From: P.Wang :      P.Wang-at-sheffield.ac.uk
Date: Wed, 31 May 2000 18:48:09 -0500
Subject: Turbo pump

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello, there,
Does anyone out here know that can we put alkylamine to the turbo pump
when carbon supporting film is being glow discharge?
Thanks.

Peiyi

Krebs Institute for Biomolecular Research
University of Sheffield
Firth Court
Western Bank
Sheffield
Yorkshire S10 2UH
United Kingdom
Tel: +44 (0)114 222 2000
Direct: +44 (0)114 222 2739
FAX: +44 (0)114 272 8697
E-mail: p.wang-at-sheffield.ac.uk






From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Wed, 31 May 2000 20:17:17 -0500
Subject: PP (coated) wires

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Briget wrote:
===========================================================
I will like to do cryomicrotomy of PP wires. My question is : can anyone
suggest what kind of epoxy is appropriate for cryomicrotomy? I used Epoxy
Mount and it was chattered during cryo process. The manufacturer confirm
to me that epoxy mount is not appropriate for cryomicrotomy. Can anynone
help me?
===========================================================
I am assuming you mean polypropylene coated wires and not polypropylene
monofilament. We have generally found that for coated wires samples, we
like to Pt coat it first (by sputtering), then embed in SPI-Pon™ 812 resin.
I would expect that any of the other "Epon® 812 substitutes", available
from the other major EM supplies firms would work just as well.

You will definitely want to use a diamond knife on this and you can vary the
hardness of the resin in way that gives you the best sections. Use a knife
angle that is not larger than 45°, the lower the temperature (usually) the
better.

Disclaimer: SPI Supplies offers for sale the resin and diamond knives
mentioned and performs this kind of cryoultramicrotomy for clients.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
===========================================




From: Shane Collins :      kshanec-at-gte.net
Date: Wed, 31 May 2000 17:15:18 -0700
Subject: RE: new developments in imaging systems?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hamamatsu recently introduced a new camera called the electron bombardment
ccd where accelerated electrons directly bombard a back thinned, peltier
cooled ccd. The electrons are emitted from a photocathode for applications
particularly in low light microscopy. With a full well capacity of 300,000
electrons, I wonder if this approach could be used in direct exposure of the
ccd to the electron beam.
Shane

K. Shane Collins
Scientific Instrument Company
805.444.4953 cell
310.568.9188 office
310.568.9189 fax


-----Original Message-----
} From: Paul Voyles [mailto:voyles-at-research.nj.nec.com]
Sent: Friday, May 26, 2000 9:11 AM
To: Microscopy-at-sparc5.microscopy.com


} You could probably expose your CCD chip directly to the electron bean --
} and buy a new chip every few hundred exposures or so! Electrons have

There is an even worse problem with exposing the CCD directly the
electron beam. The p-n junctions in the CCD chip have a certain
"well capacity" - a number of electron/hole pairs they can hold
before they saturate. Fast (keV) electrons are much more efficient
at producing electron/hole pairs than photons - so much so that a
pixel on the CCD would saturate after about 15 fast electrons hit it.
Just the square root N shot noise at that level is about 25% - larger
than typical TEM micrograph contrast of 10-20%. That's why a
phosphor or scintillator is necessary to transform the fast electrons
into photons.




Happy reading,
Paul






From: jim :      jim-at-proscitech.com.au
Date: Thu, 1 Jun 2000 13:36:52 +1000
Subject: RE: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Yes, it could be done "in theory". Somebody would need to figure out the
software and perhaps modify the hardware. Then we would find that the total
exposure of the specimen to the electron beam maybe a muliple of the film's
exposure. Afterall, an 8 sec film exposure would not amount in digital to
10x0.8, but we would require considerable time in between exposures. Since the
problems in the discussed circumstances are specimen movement and beam damage,
it seems that taking multiple exposures is a poor option.

Digital cameras are for some situation too sensitive to electron exposure.
Cutting back on electrons is no option since its the electrons that form the
image in the first instance.
Much easier in light microscopy . . . insert a neutral density filter.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com


} Hello Jim,
}
} just a couple of remarks to your email.
}
} your remark 1) What you can do of course with digital imaging (provided
} you have the necessary hardware), is to automatically collect larger
} areas by taking several images that overlap, reducing the advantage that
} film has in this area and perhaps offer other possibilities as I just
} explained in another posting as a response to Bill Tivol's posting.
}
} your remark 2) I am not sure you are not comparing apples and oranges.
} What you are saying is, that because of the "slowness" of film you need
} longer exposures, thereby averaging out the statistical noise of the
} electron, which is not the case for CCD cameras at short exposures,
} hence they appear more noisy. In essence what you are doing is to
} compare a short exposure image to a long exposure image. What you can do
} with a CCD camera is the following: You can get (perhaps) real-time dark
} field images and position your sample and/or decide if you want to
} actually take the image. Then you take a SERIES of images, let's say 10,
} each at an exposure time of 1/10 of the film exposure. This can be done
} automatically, of course. Finally, you add or average all of these
} images using a pattern recognition to align them first. The result: A
} dark field image that should be as noisy as the film image, with much
} less problems due to drift during long exposures, a higher dynamic range
} and visible immediately on the viewing screen.
}
} Michael
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 1675 Carr St., #105N
} Lakewood, CO 80215
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} } From: jim [mailto:jim-at-proscitech.com.au]
} Sent: Tuesday, May 30, 2000 7:13 PM
} To: 'William Tivol'
} Cc: microscopy-at-sparc5.microscopy.com
} Subject: RE: Film vs Digital
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Peter Bond is right in saying "The future is with digital image capture"
} if we
} were at some future date to count heads of digital versus film TEM
} users.
} That change will be for reasons of convenience and to save labour. These
} are
} powerful and valid reasons.
}
} Bill Tivol has given one set of applications where digital currently
} does not
} always measure up to film.
} Here are another couple of such applications.
} 1 When great enlargements are required film is superior. This is
} because of
} film's greater resolution, but more importantly, because its much, much
} easier
} to take images at moderate powers and highly enlarge. That way we take
} advantage of the TEM's greater depths of focus at low powers and of
} film's
} higher resolution/ image detail.
} 2 Whenever a TEM image is taken at low brightness (to avoid beam
} damage or at
} very high powers or in dark-field) relatively few electrons form the
} image and
} make that image grainy. Film is very slow and requires then a longer
} exposure,
} thus boosting the quantity of electrons used and improving the image.
} Digital
} is much more sensitive and so the exposure must be shortened. As a
} result the
} best digital camera will record, quiet faithfully the grainy, unsharp
} image.
}
} Many labs rarely or never use such applications and for them the reasons
} to
} change to digital now may be overwhelming.
} Cheers
} Jim Darley
} ProSciTech Microscopy PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} www.proscitech.com
}
} On Wednesday, May 31, 2000 5:18 AM, William Tivol
} [SMTP:tivol-at-wadsworth.org]
} wrote:
} }
} ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} -----------------------------------------------------------------------.
} }
} }
} } Peter Bond wrote:
} }
} } } The continuing discussion on whether film is better than digital,
} and
} } } whether we can directly compare their resolving ability is very
} interesting
} } } and throwing up some really useful technical stuff. But aren't we
} missing
} } } the point a bit? Surely the image quality is user defined - if you
} or your
} } } customer, are satisfied with the end product then the equipment has
} done
} } } its job. How often does one need to examine a micrograph to assess
} its
} } } limits? If are getting that close to a picture then you may be
} taking
} } } things out of context a bit and losing the whole concept.
} } }
} } } The future is with digital image capture, let's hope the price of
} the
} } } equipment comes down to something more attainable!
} } }
} }
} } Dear Pete,
} } If one is doing quantitative image processing, the better
} } resolution
} } available from film can be relevant. For example, if one wants to do
} } corellation averaging, one needs as many objects in the picture as
} possible,
} } and also as good resolution as needed--e.g., for 1 nm resolution of
} the
} } reconstruction, a pixel size of 0.25 nm times the magnification is
} necessary,
} } so the recording medium must have resolution equal to or better than
} that.
} } The pixel size of the scanner must also be this small--obviously
} irrelevant
} } for digital recording. A 5 micrometer scanning-pixel size at 20kx mag
} } gives a pixel resolution of 0.25 nm, and for 6.5 by 9 cm film, one
} will have
} } about 12,000 by 16,000 (taking the header into account) pixels of
} useful
} } information. This will give a broader area than that available at
} equal
} } resolution from any CCD chip now out there. I can assure you that I
} have
} } often had to determine the information limits in a particular
} micrograph.
} } Yours,
} } Bill Tivol
} }
} }
}





From: Chris Jeffree :      cjeffree-at-srv0.bio.ed.ac.uk
Date: Thu, 1 Jun 2000 09:52:22 +0000
Subject: SEM Stereo Image Analysis Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The electron energies used in this "intensifier" will be in the order of one
or several kV. In TEM however the energies are much higher thus one incident
electron will generate for example hundred electron-hole pairs in the CCD
thus saturating it very fast. Another factors are damage to the CCD chip and
X-Rays.

I was thinking about another approach. A chip consisting of matrix of
thermo-sensitive elements. Above each element there will be a metal block
with height equal or larger than the stop path for the electron energy used.
The whole thing will be cooled in a similar way as the CCDs. When this
assembly is exposed to the beam each block will increase its temperature
depending on the number of electrons stopped. After the exposure the matrix
is scanned and the temperature increase at each element is measured
(ofcourse before each exposure a reference image has to be taken).

The benefits:
- Very high efficiency. Almost every incident electron will contribute to
the image.
- Huge dynamic range.
- Linearity (after the temperature-signal characteristic of each element has
been calibrated)
- Narrow point spread function (maybe).

Problems:
- Difficult to manufacture (the metal blocks should be insulated from each
other)
- Maybe low sensitivity. I haven't calculated how much the temperature
increase will be (for example 5x5x30um Cu block hit by one 300 kV electron)
but I suspect it will be very small. Also It will depend on the sensitivity
of the thermo-measuring elements.
- Saturation. After each exposure one has to wait some time for the thing to
cool down again.

There are maybe other difficulties which do not come in mind now.

Hmmm now I start thinking about X-Rays. Actually the main part of the
incident energy will go into X-Rays thus making this device useless.

Rado

---------------------------------------------------------------------
Radostin Danev
Laboratory of Ultrastructure Research
National Institute for Physiological Sciences
Myodaiji-cho, Okazaki 444-8585, JAPAN
e-mail: rado-at-nips.ac.jp
---------------------------------------------------------------------
----- Original Message -----
} From: Shane Collins {kshanec-at-gte.net}
To: Paul Voyles {voyles-at-research.nj.nec.com} ;
{Microscopy-at-sparc5.microscopy.com}
Sent: Thursday, June 01, 2000 9:15 AM


AnalySIS has a module which does this
see Soft Imaging's web site at
http://www.soft-imaging.de
Chris

To: Microscopy-at-sparc5.microscopy.com
} From: Jim Ferreira {ferreira1-at-llnl.gov}


=====================================================================
DR CHRIS JEFFREE
BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
UNIVERSITY OF EDINBURGH
Daniel Rutherford Building
King's Buildings, Mayfield Road
EDINBURGH, EH9 3JH, Scotland, UK
Tel. #44 131 650 5345
FAX. #44 131 650 6563
Mobile 0410 585 401
email c.jeffree-at-ed.ac.uk
SEM / TEM bookings sem-at-ed.ac.uk
=====================================================================




From: Colin MacRae :      cmac-at-minerals.csiro.au
Date: Thu, 1 Jun 2000 07:11:59 -0500
Subject: Re: SEM Stereo Image Analysis Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In a message dated 5/31/00 2:52:24 PM, ferreira1-at-llnl.gov writes:

Anyone aware of software capable of generating depth profiles from digital
SEM stereo pair images? I am aware of the Oxford ISIS system/program, but
wondered if other companies offer similar software that would run
'stand-alone' in a PC and work with any digital files.

Jim
I have been using uMex software for the last 6 months which does
exactly
what you want. Given a stereo pair it will produce a 3D surface
reconstruction. It has a function that enables images to the two images to
be manually aligned prior to the calculation. This function when used
correctly seems to result in faster calculations.

Another other nice feature is that images can be output in VRML file
format. So you can view the images with shareware 3D packages. We use a
SGI for viewing as its rendering speed is significantly faster than a PC.

I suggest you check out the following web site.
http://www.alicona.com/en/products.htm

Regards
Colin MacRae

************************************************************************
Manager of Electron Microscopy Group (Clayton)

CSIRO
Minerals {mailto:colin.macrae-at-minerals.csiro.au}

PO Box 312, Clayton South, Ph. 61 3 9545 8800
Vic, 3169 Fax 61 3 9562 8919
AUSTRALIA

EM units WWW site http://www.minerals.csiro.au/em-unit/
*************************************************************************






From: Jeremy Sanderson :      jb_sanderson-at-yahoo.com
Date: Thu, 1 Jun 2000 06:31:32 -0700 (PDT)
Subject: Resolving power: film vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

I've been following the thread on film vs digital.
We all know that the resolving power of digital CCD
faceplates is approaching that of conventional film
emulsion, but isn't equivalent yet.
Can someone remind me of the number of mega-pixels
that a CCD will need to equate to ISO 100 print film,
ISO 64 or ISO100 slide film and the highest resolving
B&W film of all, Technical Pan at ISO 25 and ISO 100?
If anyone out there can lead me through the logic and
the maths, I'm sure others will also find it helpful.
I'll repost a summary of the replies that I get.

Regards, Jeremy Sanderson

__________________________________________________
Do You Yahoo!?
Send instant messages & get email alerts with Yahoo! Messenger.
http://im.yahoo.com/




From: Paul Voyles :      voyles-at-research.nj.nec.com
Date: Thu, 1 Jun 2000 10:34:26 -0500
Subject: RE: new developments in imaging systems?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Hamamatsu recently introduced a new camera called the electron bombardment
} ccd where accelerated electrons directly bombard a back thinned, peltier
} cooled ccd. The electrons are emitted from a photocathode for applications
} particularly in low light microscopy. With a full well capacity of 300,000
} electrons, I wonder if this approach could be used in direct exposure of the
} ccd to the electron beam.
} Shane

This certainly could be a significant improvement in CCD imaging.
Most of the width of the point spread function of current CCD camera
is due to photon spread in the scintillator, so presumably removing
the scintillator would allow digital images at resolutions very close
to the pixel size of the CCD chip.

I'm not familiar with the Hamamatsu camera, but I know that the
Gatan camera I currently use has a CCD with a full well capacity of
~500,000 electrons. In order for the Hamamatsu chip to work they
would have to find some way to reduce the electron/hole yield of the
incident fast electrons - maybe with a very thin CCD chip?

Paul Voyles


} K. Shane Collins
} Scientific Instrument Company
} 805.444.4953 cell
} 310.568.9188 office
} 310.568.9189 fax
}
}
} -----Original Message-----
} From: Paul Voyles [mailto:voyles-at-research.nj.nec.com]
} Sent: Friday, May 26, 2000 9:11 AM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: RE: new developments in imaging systems?
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Thu, 01 Jun 2000 11:27:44 -0400
Subject: Re: Resolving power: film vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jeremy Sanderson wrote:

} Dear All,
}
} I've been following the thread on film vs digital.
} We all know that the resolving power of digital CCD
} faceplates is approaching that of conventional film
} emulsion, but isn't equivalent yet.
} Can someone remind me of the number of mega-pixels
} that a CCD will need to equate to ISO 100 print film,
} ISO 64 or ISO100 slide film and the highest resolving
} B&W film of all, Technical Pan at ISO 25 and ISO 100?
} If anyone out there can lead me through the logic and
} the maths, I'm sure others will also find it helpful.
} I'll repost a summary of the replies that I get.
}
} Regards, Jeremy Sanderson
}

About twenty-five million pixels in a 35mm Kodachrome slide is the
number I have heard from several sources, none of which I can remember now.

Geoff
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************






From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Thu, 1 Jun 2000 09:23:53 -0600
Subject: RE: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



No, it's not really a problem. It's been done with low density
microscopy all the time. Granted, there are some technical aspects to be
overcome, but (and I can only speak for ourselves) we have done that on
a number of microscopes. You are of course correct, that 10 images at
0.8 seconds take longer than 8 seconds as the image has to be
transferred, etc. BUT: that's what beam blankers are for. It is pretty
straightforward to take an image at 0.8 seconds, then blank the beam
very quickly before taking the next image. That way you get pretty close
to the 8 sec total exposure. If there is no beam blanker on the
microscope, in most cases it can be added.

I am not sure what you mean by "too sensitive". The cameras are usually
constructed so that 1 electron from the beam creates between a few tenth
to a few counts (these are all statistical data, of course). The well
width divided by this sensitivity then determines, how many primary
electrons are needed to fully expose one pixel. For example, if the well
width is 50,000 electrons and the sensitivity is 1 count/electron, one
needs 50,000 primary electrons to fill the well. This translates into
roughly a 0.4% statistical error.

} From a practical standpoint: You can take images with most cameras when
the exposure meter on the microscope reads a couple of seconds without
overexposing the camera. On the other hand, you can reduce the intensity
of the beam until you see single electron events.

The one area where CCD cameras may be too sensitive is diffraction. The
normally huge intensity in the transmitted beam often leads to
saturation. In CCDs this can lead to blooming (the intensity spills over
into neighboring pixels). This can be taken care of with special chips
that have anti-blooming features, but this usually has some other
drawbacks. Again, this can also be overcome somewhat with multiple
exposures. Film behaves more civilized here, as it simply stops
responding to the electrons, but this makes film more or less useless
for quantitative measurements of diffraction patterns. I have done
diffraction with CCDs many times and though it does require some
tweaking, one can get very good results from them.

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: jim [mailto:jim-at-proscitech.com.au]
Sent: Wednesday, May 31, 2000 9:37 PM
To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'


Yes, it could be done "in theory". Somebody would need to figure out the

software and perhaps modify the hardware. Then we would find that the
total
exposure of the specimen to the electron beam maybe a muliple of the
film's
exposure. Afterall, an 8 sec film exposure would not amount in digital
to
10x0.8, but we would require considerable time in between exposures.
Since the
problems in the discussed circumstances are specimen movement and beam
damage,
it seems that taking multiple exposures is a poor option.

Digital cameras are for some situation too sensitive to electron
exposure.
Cutting back on electrons is no option since its the electrons that form
the
image in the first instance.
Much easier in light microscopy . . . insert a neutral density filter.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com






From: Matt Ervin :      mervin-at-ARL.mil
Date: Thu, 1 Jun 2000 11:37:20 -0400
Subject: Re: SEM Stereo Image Analysis Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Jim-
I saw an ad a while back for such software. I have since moved on
to other things before I had a chance to obtain the demo software. Below is the
contact information I have for the company selling the stereo image analysis software.
Since this is an edited version of the message I received several months
ago, the terms and conditions may have changed.
If you do try this software, I, and I am sure the rest of the listserve would be very
interested in hearing how well it
works, as I do have an occational need for this capability.

Here is the information (edited) that I received from the company offering the software:
now the evaluation version of MeX is available.
The test period of MeX is limited to 6 weeks.

The evaluation version is delivered with a small database and the
complete analysis tools. You can process your own images without
restrictions.

We also offer to preinclude your SEM-images into the database of the
evaluation version. You just have to send us an email and we give you
detailed information on how you should capture your images.


If you have any questions, please do not hesitate to contact us.


Kind regards, Johann Schweiger

=====================================
= Dipl.-Ing.(FH) Johann Schweiger
= Sales
= Alicona GdbR
= Koch-Sternfeldstr. 5
= D-83471 Berchtesgaden/Germany
= Tel. ++49 8652 964205
= Fax. ++49 8652 964207

I hope that you find the above information helpful.
Sincerely,
Matthew H. Ervin, Ph.D.





From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Thu, 1 Jun 2000 09:39:59 -0600
Subject: RE: SEM Stereo Image Analysis Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jim,
you may want to check out our web site for the Stereo software. It has
some images and examples of stereo evaluations from SEM images. The
'Stereo' part is not a stand-alone software, but can be combines with
our analySIS Docu software for a stand-alone application.
Michael


Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================

Hello folks
Anyone aware of software capable of generating depth profiles from
digital SEM stereo pair images? I am aware of the Oxford ISIS
system/program, but wondered if other companies offer similar software
that would run 'stand-alone' in a PC and work with any digital files.
Any help or suggestions along these lines would be most appreciated.
Thanks,
Jim









From: Steve Chapman :      protrain-at-emcourses.com
Date: Thu, 1 Jun 2000 15:24:59 +0100
Subject: Wet Specimens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi

Want to work with wet specimens?

Do not give up if you do not have an ESEM or LVSEM there may be hope yet?
Do you have a SEM with a rear manifold that connects directly to the DP and
do you have a BSE detector? Sorry but those SEM that have the pump directly
attached to the specimen chamber are no good for this procedure. AND I must
give all the credit to Viv Robinson who spawned this idea back in the 80's.

If you do have a manifold system you are in luck. Find a rubber bung that
will fit into the manifold at the rear of the specimen chamber. Drill (use
LN2) a 1/4" hole in the bung and then place it in the rear manifold. Switch
off or better still unplug your Everhart-Thornley detector (high voltage
plus poor vacuum = arcing!). Place you "wet" specimen in the microscope and
pump down. The bung will spoil the vacuum in the chamber for about 20
minutes or so and imaging with the BSE detector you will have your own LV
SEM.

To retain the moisture longer you may quench the specimen in LN2 before
putting it into the microscope. The frost will sublime away and you will be
able to watch what happens to the moisture etc.

We use this technique on all sorts of samples. A rubber bung is cheaper than
buying an LVSEM and the results, if you work quickly, are pretty good.

Try it?

Steve Chapman
Senior Consultant Protrain
For consultancy and professional training in EM world wide
Tel 44+ 1280 814774 Fax 814007
www.emcourses.com

Steve Chapman
Senior Consultant Protrain
For consultancy and professional training in EM world wide
Tel 44+ 1280 814774 Fax 814007
www.emcourses.com





From: Chris Jeffree :      cjeffree-at-srv0.bio.ed.ac.uk
Date: Thu, 1 Jun 2000 17:10:23 +0000
Subject: Re: Resolving power: film vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Is it just my problem, or isn't this all getting a little bit unfocussed!
As far as TEM is concerned, the CCD is not directly exposed to
the beam at all, so its effective resolution depends on the nature of
the system that presents the image to the CCD. The two
predominating technologies for transfer of the image to the CCD in
TEM cameras are a fibre-optic linkagage between a phosphor or
YAG scintillator and the CCD, or an optical coupling via a lens (for
example the excellent f1.2 50mm Zuiko macro lens by Olympus).
In both instances, the ultimate resolution of the system is probably
set by the electron sensor, which is the phosphor or YAG
scintillator. I suspect that fibre optic couplings probably degrade
that resolution, but I say that without reference to the facts, so
please correct me if I am mistaken. Optical coupling could in
principle project the spatial data recorded by the phosphor or YAG
to any desired magnification. It can thus be recorded by a CCD
using many pixels or few depending on the optical configuration.
So what do we mean by resolution in this context, when the
smallest object which can be imaged by a TEM can be projected
onto any desired quantity of CCD pixels?

To begin to answer Jeremy's question directly, we need to know
how much detail a Technical Pan negative can record. The figures
depend on processing technique and the test object luminance and
contrast, but the modulation transfer function figures published by
Kodak indicate that a spatial frequency in excess of 200 cycles per
mm is easily recordable. For a test object with contrast 100:1 they
quote 320 line pairs per mm. The CCD pixel spacing required to
achieve this feat would be 640 pixels per mm. That equates to a
requirement for 15360 x 23040 pixels to match the resolving power
of a 24x36mm Technical Pan exposure. That's 0.35 Giga pixels in
round numbers.

So what was that I heard about the death of silver imaging? I don't
think so. Not for a while yet.

Best wishes
Chris

} Dear All,
}
} I've been following the thread on film vs digital.
} We all know that the resolving power of digital CCD
} faceplates is approaching that of conventional film
} emulsion, but isn't equivalent yet.
} Can someone remind me of the number of mega-pixels
} that a CCD will need to equate to ISO 100 print film,
} ISO 64 or ISO100 slide film and the highest resolving
} B&W film of all, Technical Pan at ISO 25 and ISO 100?
} If anyone out there can lead me through the logic and
} the maths, I'm sure others will also find it helpful.
} I'll repost a summary of the replies that I get.
}
} Regards, Jeremy Sanderson
}
} __________________________________________________
} Do You Yahoo!?
} Send instant messages & get email alerts with Yahoo! Messenger.
} http://im.yahoo.com/
}


=====================================================================
DR CHRIS JEFFREE
BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
UNIVERSITY OF EDINBURGH
Daniel Rutherford Building
King's Buildings, Mayfield Road
EDINBURGH, EH9 3JH, Scotland, UK
Tel. #44 131 650 5345
FAX. #44 131 650 6563
Mobile 0410 585 401
email c.jeffree-at-ed.ac.uk
SEM / TEM bookings sem-at-ed.ac.uk
=====================================================================




From: Mark Armogida :      mark_armogida-at-tedpella.com
Date: Thu, 01 Jun 2000 09:31:04 -0700
Subject: Re: curing Epon/Araldite with UV?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We at Ted Pella Inc. have used our Pelco UVC2 UV Cryo Chamber to cure epoxies
but most of our UV curing has been with Acrylic resins. The literature we have
on Eponate Araldite does not show it to be UV curable but our chemist wouldn't
be surprised if it didn't accelerate the cure. The epoxy mixture should cure
overnight at 60 degrees C. We have a technical note on the use of
Epon-Araldite for embedding specimens and literature on our Cryo Chamber that
may help. I would be happy to fax them to you and/or have you talk with our
chemist.

Mark J Armogida
VP Engineering and Production
Ted Pella Inc.

Tamara Howard wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Greetings!
}
} Has anyone ever cured Epon-Araldite with UV? I was given a
} cobbled-together protocol by a non-microscopist post-doc; the protocol
} calls for a dry-ice, UV cure of Epon-Araldite. It says it can take from
} 3-7 days. I'm wondering if this is either a resin mix-up *or* the resin
} just set up all by itself (over that time frame I've seen it happen), just
} because it was a thin layer and left alone - nothing to do with the UV
} exposure. She has no ref. for this technique and isn't sure now where she
} got it. Sigh. I've only been able to find heat-cure protocols...anyone
} have any leads or thoughts on this UV thing?
}
} (Sorry about the cross-post for those of you on both of these servers)
}
} Thanks!
}
} Tamara (Planning to use the oven.....) Howard
} CSHL





From: Linda Boisselle :      THE-LAB-at-worldnet.att.net
Date: Thu, 01 Jun 2000 14:02:00 -0400
Subject: Seeking help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am seeking to speak to individuals who know about designing and
developing applications for indentification of rare cells in microscopic
biological preparations. Someone who knows how to optimize microscopy
autofocusing procedures. I am more interested in speaking to an
electrical engineer who is more into digital image processing. Can
anyone suggest what direction to take?

THE-LAB-at-att.net

Thanks.




From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Thu, 1 Jun 2000 15:14:38 -0500
Subject: TEM: looking for phage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

We are trying to help a client find and identify a bacteriophage. His
bacteria (Pasturella sp.) are plated out on agar and show well-defined clear
areas with sharp edges where the phage are supposed to be. So far, we have
taken carbon-coated grids and placed them gently onto the surface of the
clear areas, then lifted them off and negative-stained with PTA or uranyl
acetate. We have also run buffer across the clear areas, then pipetted it
onto the grids and stained it. A microbiologist who works with phage in
another lab has taken samples from the clear areas and concentrated them
down and we have stained those also.

So far we have found exactly two phage-like organisms in a total of about 10
grids. Not a stellar performance.

We figure the possibilities are: 1) the bacteria are being killed by
something other than phage; 2) we're looking for a particular type of phage
that may not be there, and we're just not seeing what's actually causing the
clear areas, or 3) for some reason we're just not getting the things
adhering to the grids, although we've used these methods successfully many
times before.

Does anyone else have any ideas that might help us out, especially on
technique? Our client is almost certain that phage are present. We just
can't find them.

Thanks in advance.

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/







From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Thu, 1 Jun 2000 12:46:37 -1000 (HST)
Subject: SEM: immunolabeling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


One of our SEM users would like to label biofilms with lectins. Three of
them. At once. I have used gold conjugated to goat anti-biotin to label
biotinylated lectins for TEM in the past, so I'm hoping to follow the same
kind of procedure. I have not done any immunolabeling for SEM, although
our FESEM has been used for such. I would be grateful for any advice! If
he wants to triple label, what sizes of gold would be useful? If he wants
to quantify the three, what kinds of controls for labeling efficiency
should we run? All hints ahd tips gratefully accepted!

Mahalo,
Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************





From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Thu, 1 Jun 2000 12:50:40 -1000 (HST)
Subject: Confocal: gold enhancement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have not generally followed any discussion on gold enhancement for light
microscopy (photons? I don't do photons), but now I need to ask for
someone what people suggest for immunogold enhancement for confocal
microscopy. We have 10nm gold left over from TEM, and it would be useful
to use it for the confocal experiment. Yes, we're being cheap!

Mahalo,
Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************





From: Gordon Couger :      gcouger-at-couger.com
Date: Thu, 1 Jun 2000 20:40:40 -0500
Subject: Re: Resolving power: film vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} From: "Chris Jeffree" {cjeffree-at-srv0.bio.ed.ac.uk}
To: "Jeremy Sanderson" {jb_sanderson-at-yahoo.com}
{snip}
} To begin to answer Jeremy's question directly, we need to know
} how much detail a Technical Pan negative can record. The figures
} depend on processing technique and the test object luminance and
} contrast, but the modulation transfer function figures published by
} Kodak indicate that a spatial frequency in excess of 200 cycles per
} mm is easily recordable. For a test object with contrast 100:1 they
} quote 320 line pairs per mm. The CCD pixel spacing required to
} achieve this feat would be 640 pixels per mm. That equates to a
} requirement for 15360 x 23040 pixels to match the resolving power
} of a 24x36mm Technical Pan exposure. That's 0.35 Giga pixels in
} round numbers.
}
At 16 bits this is per image this is 5,662,310,400 or 5.5 gig. I don't
think film is in any danger for a long time. We need at least 1 order
of magnitude for storage and 2 or 3 for processing. I remember
taking 4 days to process an image. And then work on the program and
trigger it again.

Gordon

Gordon Couger gcouger-at-couger.com

Stillwater, OK www.couger.com/gcouger
405 624-2855 GMT -6:00










From: Jenichen :      Jenichen-at-proscan.de
Date: Fri, 02 Jun 2000 11:14:47 +0200
Subject: Re: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,
This is the last call for a second post-doc vacancy at the
University of Barcelona (Spain) to work in the frame of a TMR
programme concerning UV coatings. (More details below).

Since we are expecting the Mid-Term evaluation of the project,
the starting date has been delayed to next September, so we
have extended the deadline for applications.

Any one interested please reply directly to paqui-at-el.ub.es
and/or send applications and a CV by mail before 30 June 2000.

Kind regards

F. Peir—

**************************************************************************
Laboratory: Electronic Materials and Engineering, Department of
Electronics, University of Barcelona.

Duration: 12-months, starting September 2000.



Hi Jim, some comments to Your remarks within Your text:

jim schrieb:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Yes, it could be done "in theory". Somebody would need to figure out the
} software and perhaps modify the hardware. Then we would find that the total
} exposure of the specimen to the electron beam maybe a muliple of the film's
} exposure. Afterall, an 8 sec film exposure would not amount in digital to
} 10x0.8, but we would require considerable time in between exposures.

Simply choose a CCD with higher readout performance (faster), but same quality.

} Since the
} problems in the discussed circumstances are specimen movement and beam damage,
} it seems that taking multiple exposures is a poor option.
}
} Digital cameras are for some situation too sensitive to electron exposure.

Not correct at all. The only thing is to choose the correct camera for the
application You work on. It is not complicated to make a digital system which
collects one count per incident electron to achieve the same signal to noise as in
the electron beam. This system will be less sensitive than normally sold systems
but the main advantage of digital systems that You see what You get remains and You
get instant results of Your work.
The only problem You have to solve for these systems is to use a not very sensitive
scintillator but a very high performance slow-scan CCD. So You get less visible
photons from Your electron which have to be converted in one digital count. But
these digital counts must be more than presently available to achieve a good
statistic. Thats the reason for our new 16bit CCD (dynamic up to 65536 digits) for
high performance TEM applications.

}
} Cutting back on electrons is no option since its the electrons that form the
} image in the first instance.
} Much easier in light microscopy . . . insert a neutral density filter.
} Cheers
} Jim Darley
} ProSciTech Microscopy PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} www.proscitech.com

--
Best regards / Mit freundlichen Gruessen
Dr. Frank Jenichen
Proscan elektronische Systeme GmbH
Tel.: +49 8195 999 -511 Fax: -512
mailto:Jenichen-at-proscan.de
------------------------------------------------------------
More information concerning our products
and services can be found on our website
http://www.proscan.de






From: Chris Jeffree :      cjeffree-at-srv0.bio.ed.ac.uk
Date: Fri, 2 Jun 2000 11:18:10 +0000
Subject: Re: Resolving power: film vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} About twenty-five million pixels in a 35mm Kodachrome slide is the
} number I have heard from several sources, none of which I can remember now.
}
} Geoff
Geoff
Analysing this a little further:
A 25 Mbyte image = 8.3 M pixels to cover 24x36mm or 864mm^2
this represents 98.2 pixels per mm, or 49 line pairs per mm. Pixel
size 10.2 um. Resolution about 50 line pairs per mm at best.

A 25M pixels image used to capture a 24x36mm Kodachrome
slide represents 28,935 pixels per mm^2 or 170 pixels per mm.
This is equivalent to a maximum resolution of 85 line pairs per mm,
which may be on the conservative side for Kodachrome. pixel size
= 5.88um. The image will be approx. 6120x4082 pixels, generating
a file size of approx 75Mb for a 24-bit (8+8+8bit) RGB image.

To record 120 line pairs per mm, which many top 35mm camera
lenses can achieve, a minimum of 240 pixels per mm are required,
each 4.2 um wide. This equates to 8640x5760 pixels for a
24x36mm frame = 50M-pixels or 50Mb in 8-bit greyscale, or 150Mb
in 24-bit RGB.

At 320 line pairs per mm (Technical Pan) the minimum required
640 pixels per mm is a pixel size of 1.56 um

Presumably for a light image the diffraction limited resolution is
approx 1/2 lambda which at 540nm is 0.27um.
So looking to the future of ultimate-performance CCDs, direct
recording of a diffraction limited light image projected onto the
sensor requires at the very least 3703 pixels per image mm or
13,717,421 pixels per mm^2 (greyscale 8-bit)

However, if we are doing light microscopy with an NA 1.4 x100
objective, how much resolving power do we need on CCD or film?

Data is at 0.27um resolution (lambda = 540nm). Let's round this to
0.3um. Magnification at 24x36mm film image is x100, so pixels
must be an absolute maximum of 30um wide to record the
significant data = 33.3 pixels per mm, equivalent to 800x1200
pixels to record the whole 35mm negative area. However, most
CCDs are much smaller than 35 mm frames, typically 1/3 inch. So
an 800x1200 pixel CCD at 8x12 mm, 1/3 of the linear dimensions
of a 35mm frame would use 9 pixels to record the smallest image
details. This is about right from the point of view of resolution, but
to record the whole 35mm frame we need about 2400x3600 pixels
on our CCD.

Note also that Technical Pan has (depending on the criterion used
to assess its performance) up to 10 times the resolving power
required to record all there is to see in a diffraction-limited LM
image made with a 100x NA 1.4 lens. So you can comfortably
afford to use a 60x NA 1.4 lens, thereby getting the same
resolution with a bigger field of view.

Many years ago, I took a photograph of a street scene using a
Canon 35mm SLR loaded with Kodak Recordak (I think this was a
single layer microfilm emulsion). Examined in a light microscope,
the image clearly, legibly recorded the brand-name of a child's
push chair. I tried to print this brand name using a DeVere point-
source enlarger with an image size of 20x30 inches produced with
a Schneider Componon lens, but was completely unable to
produce a legible image. The point I am making here is that the
combination of some high performance films, with high quality
lenses of the standard produced by the leading camera
manufacturers can record more detail on the film than you can
easily get back out by conventional printing. I suspect the same is
true of EM exposures.

So there is no contest - film beats CCDs for resolution hands
down. And you can process the image on the cheap. No money
goes to Intel or Microscoft, Adobe or Epson. But resolution is not
primarily what we buy CCDs for. We buy them primarily for instant
image capture in a format suitable for digital storage, digital
transmission quantitative data recording and image processing.

Chris

=====================================================================
DR CHRIS JEFFREE
BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
UNIVERSITY OF EDINBURGH
Daniel Rutherford Building
King's Buildings, Mayfield Road
EDINBURGH, EH9 3JH, Scotland, UK
Tel. #44 131 650 5345
FAX. #44 131 650 6563
Mobile 0410 585 401
email c.jeffree-at-ed.ac.uk
SEM / TEM bookings sem-at-ed.ac.uk
=====================================================================




From: jim :      jim-at-proscitech.com.au
Date: Fri, 2 Jun 2000 21:15:36 +1000
Subject: RE: Resolving power: film vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Gordon's and Chris' contribution look very bleak for digital, but the
comparison is not really fair, as we should look at the practical aspects too.
The unaided eye resolves lines 0.1mm apart, so to see the full (200 black+200
white) 400+ lines/mm that may be recorded on TEM film we would need to enlarge
over 40x. This requires an enlarger with a very wide angle lens to print small
portions of the negative and it is technically difficult to so enlarge a whole
negative since a 4" negative becomes 160" or over 4m in size. That degree of
enlargement is not fully useful since in a well exposed TEM negative at just
under 30x electron noise becomes the problem, meaning not enough electrons
contributed to the image. So enlargements beyond 30x are empty- and provide no
further information. Truly not a serious problem.
The practical part is that few people ever find it useful to enlarge more than
15x and most TEM images reproduced are barely the size of the original
negative, however, they are enlarged from a smaller portion thereof.
For these most common applications, digitals with 10+ megabytes provide
excellent image details and tonal gradation. However, that size image can only
cover the equivalent of a small 35mm negative equivalent and does not allow
high enlargement or choosing of an adjacent field.
It appears that the best of both worlds is the use of conventional TEM
negatives for archiving and scanning those as required for printing.

It should be noted that this discussion concerns TEM. Digitals in SEM are less
daunting and it is not problematical to produce excellent digitals, comparable
with film.
Cheers
Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Friday, June 02, 2000 11:41 AM, Gordon Couger [SMTP:gcouger-at-couger.com]
wrote:
}
}
} } From: "Chris Jeffree" {cjeffree-at-srv0.bio.ed.ac.uk}
} To: "Jeremy Sanderson" {jb_sanderson-at-yahoo.com}
} {snip}
} } To begin to answer Jeremy's question directly, we need to know
} } how much detail a Technical Pan negative can record. The figures
} } depend on processing technique and the test object luminance and
} } contrast, but the modulation transfer function figures published by
} } Kodak indicate that a spatial frequency in excess of 200 cycles per
} } mm is easily recordable. For a test object with contrast 100:1 they
} } quote 320 line pairs per mm. The CCD pixel spacing required to
} } achieve this feat would be 640 pixels per mm. That equates to a
} } requirement for 15360 x 23040 pixels to match the resolving power
} } of a 24x36mm Technical Pan exposure. That's 0.35 Giga pixels in
} } round numbers.
} }
} At 16 bits this is per image this is 5,662,310,400 or 5.5 gig. I don't
} think film is in any danger for a long time. We need at least 1 order
} of magnitude for storage and 2 or 3 for processing. I remember
} taking 4 days to process an image. And then work on the program and
} trigger it again.
}
} Gordon
}
} Gordon Couger gcouger-at-couger.com
}
} Stillwater, OK www.couger.com/gcouger
} 405 624-2855 GMT -6:00





From: jim :      jim-at-proscitech.com.au
Date: Fri, 2 Jun 2000 21:14:44 +1000
Subject: RE: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The world is full of possible solutions, but are they practical.?
To produce high-resolution, dark-field or any others TEM images that require
more electrons to form a clear image, Mike Bode would use multiple digital
exposures. The exposures could be layered and combined into one superior image.
This image would be made up of more pixel and is formed by more electrons and
so would be noise-free and hence could be further enlarged then otherwise
possible. Perhaps.
Beam blanking would largely save the specimen from beam damage and drift could
be compensated for by matching up the digitals. Great.
How much time is required between exposures to transfer a minimum 10mb image
per exposure? What would be the total time from focusing to the last exposure?
What about Z-drift in the interim requiring objective changes and what about
the total cost of this additional get-up. The mind boggles at a through focus
series.
When pushing the limits a piece of film seems more effective, cheaper and fa
ster.
Again, I don't doubt that there is now a large place for digital in TEM, but
its no panacea.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Friday, June 02, 2000 1:24 AM, Mike Bode [SMTP:mb-at-soft-imaging.com] wrote:
}
} No, it's not really a problem. It's been done with low density
} microscopy all the time. Granted, there are some technical aspects to be
} overcome, but (and I can only speak for ourselves) we have done that on
} a number of microscopes. You are of course correct, that 10 images at
} 0.8 seconds take longer than 8 seconds as the image has to be
} transferred, etc. BUT: that's what beam blankers are for. It is pretty
} straightforward to take an image at 0.8 seconds, then blank the beam
} very quickly before taking the next image. That way you get pretty close
} to the 8 sec total exposure. If there is no beam blanker on the
} microscope, in most cases it can be added.
}
} I am not sure what you mean by "too sensitive". The cameras are usually
} constructed so that 1 electron from the beam creates between a few tenth
} to a few counts (these are all statistical data, of course). The well
} width divided by this sensitivity then determines, how many primary
} electrons are needed to fully expose one pixel. For example, if the well
} width is 50,000 electrons and the sensitivity is 1 count/electron, one
} needs 50,000 primary electrons to fill the well. This translates into
} roughly a 0.4% statistical error.
}
} } From a practical standpoint: You can take images with most cameras when
} the exposure meter on the microscope reads a couple of seconds without
} overexposing the camera. On the other hand, you can reduce the intensity
} of the beam until you see single electron events.
}
} The one area where CCD cameras may be too sensitive is diffraction. The
} normally huge intensity in the transmitted beam often leads to
} saturation. In CCDs this can lead to blooming (the intensity spills over
} into neighboring pixels). This can be taken care of with special chips
} that have anti-blooming features, but this usually has some other
} drawbacks. Again, this can also be overcome somewhat with multiple
} exposures. Film behaves more civilized here, as it simply stops
} responding to the electrons, but this makes film more or less useless
} for quantitative measurements of diffraction patterns. I have done
} diffraction with CCDs many times and though it does require some
} tweaking, one can get very good results from them.
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 1675 Carr St., #105N
} Lakewood, CO 80215
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} From: jim [mailto:jim-at-proscitech.com.au]
} Sent: Wednesday, May 31, 2000 9:37 PM
} To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} Subject: RE: Film vs Digital
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Yes, it could be done "in theory". Somebody would need to figure out the
}
} software and perhaps modify the hardware. Then we would find that the
} total
} exposure of the specimen to the electron beam maybe a muliple of the
} film's
} exposure. Afterall, an 8 sec film exposure would not amount in digital
} to
} 10x0.8, but we would require considerable time in between exposures.
} Since the
} problems in the discussed circumstances are specimen movement and beam
} damage,
} it seems that taking multiple exposures is a poor option.
}
} Digital cameras are for some situation too sensitive to electron
} exposure.
} Cutting back on electrons is no option since its the electrons that form
} the
} image in the first instance.
} Much easier in light microscopy . . . insert a neutral density filter.
} Cheers
} Jim Darley
} ProSciTech Microscopy PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} www.proscitech.com
}





From: Horst Neve :      neve-at-bafm.de
Date: Fri, 2 Jun 2000 08:40:59 -0500
Subject: TEM - Looking for phage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Randy,

if possible, select plates obtained from serial dilutions which show
confluent lysis (i.e., where plaques touch each other). They are a good
source for obtaining high-titer bacteriophage lysates (i.g., 10E10 / 10E11
plaque-forming units per ml). Harvest the phage by scraping the top agar
and transfer it into a test tube or equivalent. Rinse the plates with some
ml of phage diluent (i.g., 5 ml). Shake this supension containing phage,
host cells and agar for some while before spinning down the cells and the
agar. It is also a good idea to pick up a single plaque. Resuspend it in a
small volume of broth, and prepare a fresh lysate in some ml of broth with
fresh host cells. For rapid screening, we sometimes pipette a drop of
} buffer onto a plaque and float a piece of carbon film into the drop
directly from a mica support. After some minutes, we pick up the film with
a grid and do the routine negative staining. But you are right, the number
of phage particles is low in this case.
Best regards
Horst Neve

} At 15:14 01.06.00 -0500, you wrote:
} Hi,
}
} We are trying to help a client find and identify a bacteriophage. His
} bacteria (Pasturella sp.) are plated out on agar and show well-defined clear
} areas with sharp edges where the phage are supposed to be. So far, we have
} taken carbon-coated grids and placed them gently onto the surface of the
} clear areas, then lifted them off and negative-stained with PTA or uranyl
} acetate. We have also run buffer across the clear areas, then pipetted it
} onto the grids and stained it. A microbiologist who works with phage in
} another lab has taken samples from the clear areas and concentrated them
} down and we have stained those also.
}
} So far we have found exactly two phage-like organisms in a total of about 10
} grids. Not a stellar performance.
}
} Thanks in advance.
}
} Randy
}
} Randy Tindall
} EM Specialist
} Electron Microscopy Core Facility
} W122 Veterinary Medicine
} University of Missouri
} Columbia, MO 65211

****************************************************************************
****
Dr. Horst Neve
Institut fuer Mikrobiologie / Institute for Microbiology
Bundesanstalt fuer Milchforschung / Federal Dairy Research Centre
Postfach / P.O. Box 6069, D-24121 Kiel
Hermann-Weigmann-Str. 1, D-24103 Kiel
****************************************************************************
****
Tel. / Phone: +49 (0) 431 609 2343 {} Fax: +49 (0) 431 609 2306
E-mail: neve-at-bafm.de {} Internet: http://www.bafm.de
****************************************************************************
****






From: Horst Neve :      neve-at-bafm.de
Date: Fri, 2 Jun 2000 08:42:08 -0500
Subject: TEM - Looking for phage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Randy,

if possible, select plates obtained from serial dilutions which show
confluent lysis (i.e., where plaques touch each other). They are a good
source for obtaining high-titer bacteriophage lysates (i.g., 10E10 / 10E11
plaque-forming units per ml). Harvest the phage by scraping the top agar
and transfer it into a test tube or equivalent. Rinse the plates with some
ml of phage diluent (i.g., 5 ml). Shake this supension containing phage,
host cells and agar for some while before spinning down the cells and the
agar. It is also a good idea to pick up a single plaque. Resuspend it in a
small volume of broth, and prepare a fresh lysate in some ml of broth with
fresh host cells. For rapid screening, we sometimes pipette a drop of
} buffer onto a plaque and float a piece of carbon film into the drop
directly from a mica support. After some minutes, we pick up the film with
a grid and do the routine negative staining. But you are right, the number
of phage particles is low in this case.
Best regards
Horst Neve

} At 15:14 01.06.00 -0500, you wrote:
} Hi,
}
} We are trying to help a client find and identify a bacteriophage. His
} bacteria (Pasturella sp.) are plated out on agar and show well-defined clear
} areas with sharp edges where the phage are supposed to be. So far, we have
} taken carbon-coated grids and placed them gently onto the surface of the
} clear areas, then lifted them off and negative-stained with PTA or uranyl
} acetate. We have also run buffer across the clear areas, then pipetted it
} onto the grids and stained it. A microbiologist who works with phage in
} another lab has taken samples from the clear areas and concentrated them
} down and we have stained those also.
}
} So far we have found exactly two phage-like organisms in a total of about 10
} grids. Not a stellar performance.
}
} Thanks in advance.
}
} Randy
}
} Randy Tindall
} EM Specialist
} Electron Microscopy Core Facility
} W122 Veterinary Medicine
} University of Missouri
} Columbia, MO 65211

****************************************************************************
****
Dr. Horst Neve
Institut fuer Mikrobiologie / Institute for Microbiology
Bundesanstalt fuer Milchforschung / Federal Dairy Research Centre
Postfach / P.O. Box 6069, D-24121 Kiel
Hermann-Weigmann-Str. 1, D-24103 Kiel
****************************************************************************
****
Tel. / Phone: +49 (0) 431 609 2343 {} Fax: +49 (0) 431 609 2306
E-mail: neve-at-bafm.de {} Internet: http://www.bafm.de
****************************************************************************
****






From: Phillip Rutledge :      prutledge-at-ars.usda.gov
Date: Fri, 02 Jun 2000 08:23:17 -0600
Subject: Incubators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microbiologist experts:

I'm looking for a (preferably) bench top incubator. Non-water jacketed, not using B.O.D. bottles, unit needs to be using CFC free refrigeration system. Does anyone know of such a unit or where one can be purchased? Need to order one ASAP. We are a small research lab and the one we have is a monster (48"Hx46"Wx28"deep and weighs almost 500lbs.). I'd appreciate any info.

Thanks,

Phil Rutledge
prutledge-at-ars.usda.gov




From: Beverly_E_Maleeff-at-sbphrd.com
Date: Fri, 2 Jun 2000 10:50:05 -0400
Subject: M&M2000: Social Event Information

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Microscopy & Microanalysis 2000 Local Arrangements Committee
is pleased to announce an additional social event at M&M2000---

TAKE ME OUT TO THE BALL GAME !!

Join us on Tuesday, August 15th as the
PHILADELPHIA PHILLIES meet the
ARIZONA DIAMONDBACKS at Veterans Stadium.
Game time is 7:35 PM.

For $32.50, you'll get:
- a ticket to the game (300 level, on the third base line)
- a $10.00 coupon, good toward purchases at all concession and souvenir stands
- round-trip transportation from the Convention Center to the stadium
on a luxury coach (buses leave at 6:00 PM; return to the convention
center immediately following the game)

A limited number of packages are still available. First come, first served.

Checks, in US$ only, payable to "Microscopy & Microanalysis 2000" can be mailed
to:

Bev Maleeff
Treasurer, M&M2000 LAC
c/o SmithKline Beecham Pharmaceuticals
Mail Code UE 0462
709 Swedeland Road
King of Prussia, PA 19406

Checks only, please. No credit cards accepted.
Your cancelled check will serve as your receipt.

Tickets will be distributed at the M&M2000 Hospitality Booth
on Monday and Tuesday, August 14th & 15th.

Don't miss out on the fun!!!

Further information about this and other M&M2000 events can be found on our
website:
http://www.msa.microscopy.com/~mm2000/

See you in August!!

Bev Maleeff
M&M2000 LAC






From: William McManus :      billemac-at-biology.usu.edu
Date: Fri, 2 Jun 2000 09:16:34 -0600
Subject: TEM: Freeze Fracture help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Date sent: Fri, 02 Jun 2000 11:03:03 -0400
To: c.jeffree-at-ed.ac.uk
} From: joe fu {jofu-at-nist.gov}


I have a researcher here at USU who would like to have some freeze
fracture preps made of lysosome. Is there anyone out there who is
currently doing FF preps and would be willing to assist us? I can image
the preps here in Logan.

William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920





From: John J. Bozzola :      bozzola-at-siu.edu
Date: Fri, 2 Jun 2000 10:43:42 -0500
Subject: Re: TEM: looking for phage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Rando,

Having worked with a variety of phages (Candida, Streptococcus, E.
coli) I can tell you that finding phages from an agar plaque is very
difficult--as you have determined. The best way is to do a liquid
culture and then high speed followed by ultracentrifugation to
concentrate the particles. FYI, as I recall, on a 200 mesh grid each
virus particle is roughly equivalent to 3.4 x 10E6 vp/ml. So, you
need a lot of particles to even find one of them using this approach.

JB
####################################################################
John J. Bozzola, Ph.D., Director
Micro-Imaging and Analysis Center
750 Communications Drive - MC 4402
Southern Illinois University
Carbondale, IL 62901 U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################




From: Ruth Yamawaki :      yamawaki-at-stanford.edu
Date: Fri, 2 Jun 2000 09:48:03 -0700
Subject: Outside electrical near the EM lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My EM lab is going to be moved and I have just seen the plans. Apparently
two large equipment room are going to be built, one right across the hall
and the other two doors down. I have been told that the equipment room will
house -80 freezers, centrifuges, etc. Does anyone out there have experience
with this type of equipment near their TEM? I will not be able to test the
room before the move because nothing has been built yet and we all scheduled
to move in at the same time. I am hoping to have some influence on the
architects now. They have been told (and I shall keep reminding them) that
no electrical circuits are to be passed around the EM room.

Thanks.

Ruth

**********************************************
Ruth Yamawaki
Department of Comparative Medicine
Building 330, Quad 7, RAF-1
Stanford CA 94305
(650) 723-3457
**********************************************





From: Fortner, Jeffrey A. :      fortner-at-cmt.anl.gov
Date: Fri, 2 Jun 2000 14:36:59 -0500
Subject: RE: Outside electrical near the EM lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Well, let's see:

Jim wrote:

Mike Bode would use multiple digital
exposures. The exposures could be layered and combined into one superior
image.
This image would be made up of more pixel and is formed by more
electrons and
so would be noise-free and hence could be further enlarged then
otherwise
possible. Perhaps.

No, I did not talk about further enlargements. All I wanted to say is,
that a more noise-free image can be achieved by adding multiple images,
and that this also to some extent helps with drift of the sample during
acquisition.

Jim wrote:

How much time is required between exposures to transfer a minimum 10mb
image
per exposure?

How did you arrive at 10 MB? A 1280x1024 image with 16 bit pixel
information is about 2.5 MB (uncompressed). We acquire about 10 of those
per second and transfer them across the PC bus to the display. Putting
them on them into Memory might add a few tenth of a second. Writing to
HD can be done after all images are acquired.

Jim wrote:

What would be the total time from focusing to the last exposure?
What about Z-drift in the interim requiring objective changes

Why would we have to worry about that, if we don't have to worry about
that when taking the image on film? In fact, we could take care of this
by looking at the image between exposures and correct for z-drift.
However, as you said, that would add to the overall time and exposure. I
was comparing a normal dark field image taken on film at 8 seconds with
acquiring the same image on a "too sensitive" CCD camera by adding up 10
consecutive .8 second images. Why would the sample drift (in x, y or z)
substantially more in 8+delta seconds than in 8?

Jim wrote:

what about the total cost of this additional get-up

That of course depends on the microscope and there is no general answer.
For example on a LEO 912 I believe the blanker is standard. The
additional cost to use an acquisition scheme like this with our software
is $0 plus perhaps a bit of time to write a small macro. On other
microscopes one might have to add a beam blanker and perhaps a control
mechanism for the beam blanker. But I would guess, that this cost is not
very high. All modern microscopes are computer controller anyway, so it
is most likely just a control command that needs to be sent to the
microscope over a serial port if the beam blanker is installed. Piece of
cake.

Jim wrote:

The mind boggles at a through focus series.

You're right here. But I don't think we were talking about through-focus
series. Incidentally, we do through-focus series on light microscopes
and reconstruction routinely. Takes a few images at different focus (or
for a light microscope: stage) settings. The rest is done off-line.
Takes maybe a couple of minutes for about 20 images of about 1kx1k. I
agree that TEM is different here and much more complicated due to the
complicated Contrast Transfer Function. However, this could in principle
be sorted out.

Jim wrote:

Again, I don't doubt that there is now a large place for digital in TEM,
but
its no panacea.

I also agree with you on that one. But using the additional computer
possibilities of digital imaging might take you further than expected.

Michael


Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: jim [mailto:jim-at-proscitech.com.au]
Sent: Friday, June 02, 2000 5:15 AM
To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
Cc: 'jim-at-proscitech.com.au'


This can be hard to predict. Once we were having trouble with rather huge
(100 eV) energy fluctuations in our GIF 200 energy filter. We traced the
source to an adjoining room filled with constant-temperature ovens, fans,
and other high-current equipment. The unlikely source finally turned out to
be a $50 hot plate-stirrer!

Keeping the AC circuitry from running near the lab is a very good
start- I know this has devastated other labs. Ask for your own independent
electrical ground for the 'scope, and that all electrical circuits to your
lab remain independent of other rooms. I doubt that the equipment you
describe will be a big problem as long as you don't share AC circuits,
ground, or a common wall.
Good luck.


"The chief source of problems is solutions."
-Eric Sevareid

...........................................................................
......................................................
Jeffrey A. Fortner
Argonne National Laboratory
CMT/205
9700 S. Cass Avenue
Argonne, IL 60439-4837

(630) 252-5594 (voice)
(630) 252-4771 (fax)





} ----------
} From: Ruth Yamawaki
} Sent: Friday, June 2, 2000 11:48 AM
} To: 'Microscopy-at-sparc5.microscopy.com'
} Subject: Outside electrical near the EM lab
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} My EM lab is going to be moved and I have just seen the plans. Apparently
} two large equipment room are going to be built, one right across the hall
} and the other two doors down. I have been told that the equipment room
} will
} house -80 freezers, centrifuges, etc. Does anyone out there have
} experience
} with this type of equipment near their TEM? I will not be able to test
} the
} room before the move because nothing has been built yet and we all
} scheduled
} to move in at the same time. I am hoping to have some influence on the
} architects now. They have been told (and I shall keep reminding them)
} that
} no electrical circuits are to be passed around the EM room.
}
} Thanks.
}
} Ruth
}
} **********************************************
} Ruth Yamawaki
} Department of Comparative Medicine
} Building 330, Quad 7, RAF-1
} Stanford CA 94305
} (650) 723-3457
} **********************************************
}
}




From: Suzannah Mayo :      smayo-at-southernmicro.com
Date: Fri, 02 Jun 2000 17:02:06 -0400
Subject: Re: TEM - Looking for phage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There is a standard technique for phage isolation and purification in Maniatis.
I have found it to work better than any kit and can be modified accordingly.

Horst Neve wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Randy,
}
} if possible, select plates obtained from serial dilutions which show
} confluent lysis (i.e., where plaques touch each other). They are a good
} source for obtaining high-titer bacteriophage lysates (i.g., 10E10 / 10E11
} plaque-forming units per ml). Harvest the phage by scraping the top agar
} and transfer it into a test tube or equivalent. Rinse the plates with some
} ml of phage diluent (i.g., 5 ml). Shake this supension containing phage,
} host cells and agar for some while before spinning down the cells and the
} agar. It is also a good idea to pick up a single plaque. Resuspend it in a
} small volume of broth, and prepare a fresh lysate in some ml of broth with
} fresh host cells. For rapid screening, we sometimes pipette a drop of
} } buffer onto a plaque and float a piece of carbon film into the drop
} directly from a mica support. After some minutes, we pick up the film with
} a grid and do the routine negative staining. But you are right, the number
} of phage particles is low in this case.
} Best regards
} Horst Neve
}
} } At 15:14 01.06.00 -0500, you wrote:
} } Hi,
} }
} } We are trying to help a client find and identify a bacteriophage. His
} } bacteria (Pasturella sp.) are plated out on agar and show well-defined clear
} } areas with sharp edges where the phage are supposed to be. So far, we have
} } taken carbon-coated grids and placed them gently onto the surface of the
} } clear areas, then lifted them off and negative-stained with PTA or uranyl
} } acetate. We have also run buffer across the clear areas, then pipetted it
} } onto the grids and stained it. A microbiologist who works with phage in
} } another lab has taken samples from the clear areas and concentrated them
} } down and we have stained those also.
} }
} } So far we have found exactly two phage-like organisms in a total of about 10
} } grids. Not a stellar performance.
} }
} } Thanks in advance.
} }
} } Randy
} }
} } Randy Tindall
} } EM Specialist
} } Electron Microscopy Core Facility
} } W122 Veterinary Medicine
} } University of Missouri
} } Columbia, MO 65211
}
} ****************************************************************************
} ****
} Dr. Horst Neve
} Institut fuer Mikrobiologie / Institute for Microbiology
} Bundesanstalt fuer Milchforschung / Federal Dairy Research Centre
} Postfach / P.O. Box 6069, D-24121 Kiel
} Hermann-Weigmann-Str. 1, D-24103 Kiel
} ****************************************************************************
} ****
} Tel. / Phone: +49 (0) 431 609 2343 {} Fax: +49 (0) 431 609 2306
} E-mail: neve-at-bafm.de {} Internet: http://www.bafm.de
} ****************************************************************************
} ****





From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Fri, 02 Jun 2000 14:43:49 -0700
Subject: Fwd: RE: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Jim

I spent about two years trying to make good pictures of my NanoGold labeled
protein-DNA complexes. Doing this job I find two main problems: the
sample is unstable under the beam as any biological sample; NanoGold is
much more bright than protein core in dark-field. I find that it is
impossible to record equally perfect signals from NanoGold and protein core
because of short dynamic range for SO-163 film, I believe. I was trying to
make two pictures with different exposure, but it is tricky: in dark-field
mode the automatic exposure meter usually does not work and we have to set
exposure manually, in this case it is difficult to get "right" exposure
time in the right moment, you know. Again, because of sample's short life
under the beam, it is impossible to make a couple pictures at the different
conditions sometime. Keep in mind, please, that to change the film in the
microscope it takes about 10 seconds. Your idea about increasing
signal-noise ratio by collecting more electrons is bright but not
practical. For biological samples (I am talking about non-fixed,
non-stained samples of proteins, DNA or RNA-protein complexes etc) the
electron damage is a huge problem. People are trying to solve it in
different ways. Some using cryo temperature (to stabilize the biological
structure). I was using freeze-drying (I find that freeze-dried samples
are more stable under the beam). But in any case we have deal with very
unstable samples and must to do everything to decrease (not increase as you
recommended) electron dose. Drift is a second big problem for such
application: to increase signal-noise ratio we have to use very thin
support films. Images obtained at such conditions are noisy and in most
cases we have to use image analysis tools to extract the data. It means
that we have to digitize our images anyway. In such situation digital
camera may help. As you, probably, remember I was a person who initiates
this discussion. I think this discussion was very useful for many of us
who are not friendly with digital camera's techniques. We understand the
limitations of the modern digital cameras better now. I would like to say
thank you everybody who was involved in this discussion. There is some
conclusions I make for myself from discussion:


- Film is still cheap and universal material for recording and
storage EM images, sorry CCD.
- CCD TEM camera should not substitute film. Film and camera should
work all together improving the flexibility of the TEM system. For this
reason I will chose side-mount camera if will have money for it.
- For cell-biology (thin sections) where the resolution of the sample
is about 3 nm CCD camera may do a good job allowing users to make a huge
number of pictures (cell-biology guys love it), instantly view and
catalogize them.
- Sometime the digital camera may help in area of high-resolution
(relatively high, guys) EM when image will be digitized anyway. The major
limitation here is small area of view (we need a lot of particles for image
analysis sometime), but you could make the set of overlapping pictures and
digitally combine it. I love, also, Mike Bode idea to make a few very
short exposure pictures and combine it digitally later to reduce noise. The
relatively big size of CCD's pixels is a real problem too.
- CCD camera is expensive "toy". I am not sure that the benefits from
using it will compensate astronomical price, actually the third of the
electron microscope value ($70000 is it 1/3 of microscope's price on
current market?). Currently, I would recognize the CCD TEM camera as funny
"attachment" which may be useful if you rich enough to spend money on it
(it mean, that you have everything else in your EM lab plus some extra
$70000 for the fun playing with digital "toy").
- TEM CCD cameras are under extensive development now. Today's
camera will be replaced on the new model (read better, faster, what else?)
next year. Next year's camera will be easily forgotten next after the next
year and so on... Each new camera will be better that previous one... CCD
TEM camera it is not a good investment of money, I think.
- We should keep in mind that many companies charged extra 3-4K$ for
the installation and training (it is mandatory for Gatan for instance) and
you, probably, have to buy service contract on it even if you have service
contract on the microscope (JEOL's service contract on microscope do not
cover the CCD camera even if you buy it from JEOL).

Best regard, Sergey




} Date: Fri, 02 Jun 2000 21:14:44 +1000
} From: jim {jim-at-proscitech.com.au}
} Subject: RE: Film vs Digital
} To: 'Mike Bode' {mb-at-Soft-Imaging.com} ,
} "'Microscopy-at-MSA.Microscopy.Com'" {Microscopy-at-sparc5.microscopy.com}
} Cc: "'jim-at-proscitech.com.au'" {jim-at-proscitech.com.au}
} Reply-to: "jim-at-proscitech.com.au" {jim-at-proscitech.com.au}
} Organization: ProSciTech
} X-Mailer: Microsoft Internet E-mail/MAPI - 8.0.0.4211
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
Pager: (310) 845-0248
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu
http://www.bol.ucla.edu/~sryazant






From: Jonathan Barnard :      Jonathan.Barnard-at-angstrom.uu.se
Date: Sat, 03 Jun 2000 14:19:19 +0200
Subject: TEM:beam damage vs voltage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chris,

I agree with most of your statements, and I don't think that anybody
would argue the point, that a raw CCD chip has a better resolution that
film. As you pointed out, film can have a very small grain size ( {1
micron) and CCD chips usually have a few microns pixel size.

But that is not the end of the story. An optical system normally
consists of more than a chip or a sheet of film. The question is, can I
get the resolution I want or need. And here the situation is not as
simple. For example: I used to do high-resolution TEM. What you do there
is operate the microscope at optimum condition, then take a picture (on
film). You then go to the darkroom and develop prints by blowing up the
negative 10, 20 or even more times. When you then look at the images,
you can usually see the grains of the film (especially if you then scan
those into a copmputer). So, we are working at the resolution limit of
the film, and according to your postings, we should not be able to see
anything on a CCD. But that's not true. By using some geometrical
properties (the camera sits further down in the column and sees an
already enlarged image) and a tapered fiber-optic, you can acquire just
as good and better images of the same structure. Both images are limited
by the point resolution of the TEM and not by the Film or CCD
resolution.

What I am trying to say, and what I have said before is, that film is
definitely better when it comes to maximizing the product of resolution
AND field of view. However, if we can trade one for the other, I believe
in most cases you get better results from a CCD.

Having said that and looking at a print of Ansel Adams, I am glad there
is film, though!!

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Chris Jeffree [mailto:cjeffree-at-srv0.bio.ed.ac.uk]
Sent: Friday, June 02, 2000 10:16 AM
To: joe fu
Cc: microscopy-at-sparc5.microscopy.com


Date sent: Fri, 02 Jun 2000 11:03:03 -0400
To: c.jeffree-at-ed.ac.uk
} From: joe fu {jofu-at-nist.gov}


Conventional wisdom said that when it came to using TEM for biological
specimens (or beam sensitive specimens) it is always best to use a lower
accelerating voltage. Many lower voltage microscopes (V { 120 kV) were sold
on this assumption.
However is this assumption true? At the present there are many research
establishments, buying TEMs for biological use, who are using 200 to 300 kV
beams. So obviously there has been a shift in the conventional way of
thinking. Being materials based I am not sure what the status quo is in
biological TEM.
I know that the ratio of inelastic to elastic scattering cross sections is
greater than one for the elements Z {12, but how does this change as the
beam energy increases? What are your experiences? This is really just
academic curiosity by the way, but I am sure many of you would appreciate
the question.
********************************************************
Dr Jonathan Barnard

Analytical Materials Physics
The Angstrom Laboratory, Uppsala University
P O Box 534, SE-751 21 Uppsala, Sweden
Phone: +46-(0)18-4716838 Fax: +46-(0)18-500131
Phone: Microscope room +46 18 471 6365
http://www.angstrom.uu.se/analytical/home.html
********************************************************





From: Larry Allard :      l2a-at-ornl.gov
Date: Sat, 03 Jun 2000 11:46:51 -0400
Subject: Sergey's comments Re: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sergey writes:

} } - CCD camera is expensive "toy". I am not sure that the
} } benefits from using it will compensate astronomical price, actually
} } the third of the electron microscope value ($70000 is it 1/3 of
} } microscope's price on current market?). Currently, I would
} } recognize the CCD TEM camera as funny "attachment" which may be
} } useful if you rich enough to spend money on it (it mean, that you
} } have everything else in your EM lab plus some extra $70000 for the
} } fun playing with digital "toy").
- TEM CCD cameras are under extensive development now. Today's
camera will be replaced on the new model (read better, faster, what
else?) next year. Next year's camera will be easily forgotten next
after the next year and so on... Each new camera will be better that
previous one... CCD TEM camera it is not a good investment of money,
I think. { {


Sergey:

I appreciate your post, but...


1. Remember that "time is money"...there is no question that there
is a value in the nearly instantaneous result that derives from the
use of digital imaging systems, particular on TEMs. Also, the
ability to rapidly process and analyze your images to determine if
they are "keepers" is priceless, IMO.

2. In our large, national multi-user facility we have been
digital-only since about 1993 or so. We have no darkrooms for plate
loading or enlarging. The only "chemicals" we handle in the entire
photographic process are toner cartridges for our laser printers, or
equivalent media for dye sub printers etc. In most instances, we
process our images electronically all the way through to the final
presentation. This includes preparation of PowerPoint slides for
digital projection for talks, to sending full papers out for
publication on disks or via e-mail attachments. No user has *ever*
complained about the non-availability of film, or about the "lack of
resolution of CCD images" compared to film. On the contrary, we have
users that travel to our laboratory specifically to do work on our
instruments because of the availability of digital imaging, when the
same instruments in their own laboratories are not equipped with
digital cameras.

3. Digital camera systems also provide the capability to conduct
research with outside colleagues live-time via Telepresence
Microscopy methods. This is a rapidly developing capability in our
field that just cannot be done (on a TEM at least) if your
microsocope only uses film for imaging.

4. The 1k x 1k Gatan CCD camera on our Hitachi HF-2000 was purchased
for about $100K in 1993, and was upgraded about 5 years ago to a
multi-scan capability. It is still functioning perfectly today, as
is a similar camera on our JEOL 4000EX TEM. The very newest CCD
cameras might offer faster read-out times, but there has been no
order-of-magnitude improvement in capabilities to make our older
camera so obsolete that we desire to purchase a new one. The point
is that, unlike desk-top computers, the *expensive* digital camera
you purchase today will definitely *not* be obsolete next year...

5. A CCD TEM camera is probably the *best* investment anyone could
make to advance the research throughput in their microscopy facility.

All just MHO...

Larry

Dr. Lawrence F. Allard
Senior Research Staff Member
High Temperature Materials Laboratory
Oak Ridge National Laboratory
1 Bethel Valley Road
Bldg. 4515, MS 6064
PO Box 2008
Oak Ridge, TN 37831-6064

865-574-4981
865-576-5413 Fax
allardlfjr-at-ornl.gov




From: Gordon Couger :      gcouger-at-couger.com
Date: Sat, 3 Jun 2000 12:53:11 -0500
Subject: Re: Resolving power: film vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} From: "Mike Bode" {mb-at-Soft-Imaging.com}
To: "'Microscopy-at-MSA.Microscopy.Com'" {Microscopy-at-sparc5.microscopy.com}
} Chris,
}
} I agree with most of your statements, and I don't think that anybody
} would argue the point, that a raw CCD chip has a better resolution that
} film. As you pointed out, film can have a very small grain size ( {1
} micron) and CCD chips usually have a few microns pixel size.
}
} But that is not the end of the story. An optical system normally
} consists of more than a chip or a sheet of film. The question is, can I
} get the resolution I want or need. And here the situation is not as
} simple. For example: I used to do high-resolution TEM. What you do there
} is operate the microscope at optimum condition, then take a picture (on
} film). You then go to the darkroom and develop prints by blowing up the
} negative 10, 20 or even more times. When you then look at the images,
} you can usually see the grains of the film (especially if you then scan
} those into a computer). So, we are working at the resolution limit of
} the film, and according to your postings, we should not be able to see
} anything on a CCD. But that's not true. By using some geometrical
} properties (the camera sits further down in the column and sees an
} already enlarged image) and a tapered fiber-optic, you can acquire just
} as good and better images of the same structure. Both images are limited
} by the point resolution of the TEM and not by the Film or CCD
} resolution.c


For a scanning device be it light, electrons, x-rays or gamma rays is a CCD
array the best way to capture the image. My experience has been with gamma
rays and to some extent x rays. We have not found a good way to focus gamma
rays so we use a crystal that emitted light and a photomultiplier tube and
we
got good images. The resolution depended on the aperture of the gamma ray
source and was pretty large.

If you are scanning a sample wiht an electron beam the same principle should
work. You would not need long dwell times but build the images out of
multiple
scans.

If you are scanning the sample the array of pixels seems redundant to me.
Also a photomultiplier tube and crystal are a great deal more sensitive than
CCD arrays and have a good deal more dynamic range.

The time to make a film image is allows going to be less than making a
digital image.
The ability to immediately see the digital image is a very handy thing.
There are ways
to develop B&W film that you can see your image in 5 minutes. For 4 X 5
images I am
using BZT tubes that are tubs with a inter circumference of a little over 4
inches. You
put the file and developer in the tube and put the tub in a pan of water and
spin the tube
in the water. The stop bath and fixing can be carried out in room light and
it results in
the most even development I have ever seen.

For archival storage I don't think 35 mm film can be beat for cost and
resolution. Unfortunately
you can't see the results until later occasional making it necessary to
reshoot the session if
you rely on film alone.

If you really need long term archival storage you might consider using a
slide printer
for a computer to print the digital images. We know the life of silver film
is longer then
100 years. For the LM folks that are using color images you could use
PhotoShop to
do a color separation and print out three negatives on black and white for
storage.

Hard drives And CD ROMS have come a long way but I loose about 20% of my
hard
drive storage a year and loose an occasional CDROM to scratches. Film would
survive
the scratches with a minimal loss of information insted loosing the whole
CDROM.

Gordon

Gordon Couger gcouger-at-couger.com

Stillwater, OK www.couger.com/gcouger
405 624-2855 GMT -6:00








From: STANSMAN-at-aol.com
Date: Sat, 3 Jun 2000 17:21:51 EDT
Subject: Re: Resolving power: film vs digital Commercial Post

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

Please be aware that this is a commercial post that may be of interest to
some of the list members.


Digital light microscope cameras are now available that use a highly touted
CCD detector: Color or monochrome; fast interline transfer, high sensitivity,
low noise; 6.7 micron square pixel; 2/3 inch 1.3 Million chip; PCI bus frame
transfer. This detector is ideal to use in light microscopy applications.
Now precision technology and fast high capacity cheap computers allow this
detector to be "Micro Stepped" providing digital images of up to 12 million
pixels to be captured very quickly and providing file sizes of up to 35 MB in
24 bit color images. "Inter Pixel Stepping" technology will allow a
considerably less expensive approach to digital imaging that can now approach
film resolving capability in a light microscope.

Hats of to Dr. Jeffree's explanation on optical resolution as it relates to
light microscopy applications. With high pixel density capable digital
cameras, now lower magnification, low NA (lower resolving power) objectives
can be used to acquire matching digital resolution images of wide fields of
view.

Information on Nikon's Instrument Division "Digital Eclipse" family of
digital cameras for microscopy including the new DXM1200 digital color camera
will soon be available on our web site www.nikonusa.com or you can go to
our Dealer Locator at
http://www.nikonusa.com/corpinfo/dealers/dealerSearch.cfm and ask your
local Nikon authorized microscope dealer for further information.

Best Regards,

Stan Schwartz
Manager, BioSciences Dept.
Nikon Inc Instrument Division
1300 Walt Whitman Rd.
Melville, NY 11747
631-547-8500
631-547-4033 Fax
Schwartz-at-nikonincmail.com
www.nikonusa.com


Earlier post
=============================================================
Geoff
Analysing this a little further:
A 25 Mbyte image = 8.3 M pixels to cover 24x36mm or 864mm^2
this represents 98.2 pixels per mm, or 49 line pairs per mm. Pixel
size 10.2 um. Resolution about 50 line pairs per mm at best.

A 25M pixels image used to capture a 24x36mm Kodachrome
slide represents 28,935 pixels per mm^2 or 170 pixels per mm.
This is equivalent to a maximum resolution of 85 line pairs per mm,
which may be on the conservative side for Kodachrome. pixel size
= 5.88um. The image will be approx. 6120x4082 pixels, generating
a file size of approx 75Mb for a 24-bit (8+8+8bit) RGB image.

To record 120 line pairs per mm, which many top 35mm camera
lenses can achieve, a minimum of 240 pixels per mm are required,
each 4.2 um wide. This equates to 8640x5760 pixels for a
24x36mm frame = 50M-pixels or 50Mb in 8-bit greyscale, or 150Mb
in 24-bit RGB.

At 320 line pairs per mm (Technical Pan) the minimum required
640 pixels per mm is a pixel size of 1.56 um

Presumably for a light image the diffraction limited resolution is
approx 1/2 lambda which at 540nm is 0.27um.
So looking to the future of ultimate-performance CCDs, direct
recording of a diffraction limited light image projected onto the
sensor requires at the very least 3703 pixels per image mm or
13,717,421 pixels per mm^2 (greyscale 8-bit)

However, if we are doing light microscopy with an NA 1.4 x100
objective, how much resolving power do we need on CCD or film?

Data is at 0.27um resolution (lambda = 540nm). Let's round this to
0.3um. Magnification at 24x36mm film image is x100, so pixels
must be an absolute maximum of 30um wide to record the
significant data = 33.3 pixels per mm, equivalent to 800x1200
pixels to record the whole 35mm negative area. However, most
CCDs are much smaller than 35 mm frames, typically 1/3 inch. So
an 800x1200 pixel CCD at 8x12 mm, 1/3 of the linear dimensions
of a 35mm frame would use 9 pixels to record the smallest image
details. This is about right from the point of view of resolution, but
to record the whole 35mm frame we need about 2400x3600 pixels
on our CCD.

Note also that Technical Pan has (depending on the criterion used
to assess its performance) up to 10 times the resolving power
required to record all there is to see in a diffraction-limited LM
image made with a 100x NA 1.4 lens. So you can comfortably
afford to use a 60x NA 1.4 lens, thereby getting the same
resolution with a bigger field of view.

Many years ago, I took a photograph of a street scene using a
Canon 35mm SLR loaded with Kodak Recordak (I think this was a
single layer microfilm emulsion). Examined in a light microscope,
the image clearly, legibly recorded the brand-name of a child's
push chair. I tried to print this brand name using a DeVere point-
source enlarger with an image size of 20x30 inches produced with
a Schneider Componon lens, but was completely unable to
produce a legible image. The point I am making here is that the
combination of some high performance films, with high quality
lenses of the standard produced by the leading camera
manufacturers can record more detail on the film than you can
easily get back out by conventional printing. I suspect the same is
true of EM exposures.

So there is no contest - film beats CCDs for resolution hands
down. And you can process the image on the cheap. No money
goes to Intel or Microscoft, Adobe or Epson. But resolution is not
primarily what we buy CCDs for. We buy them primarily for instant
image capture in a format suitable for digital storage, digital
transmission quantitative data recording and image processing.

Chris

=====================================================================
DR CHRIS JEFFREE
BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
UNIVERSITY OF EDINBURGH
Daniel Rutherford Building
King's Buildings, Mayfield Road
EDINBURGH, EH9 3JH, Scotland, UK
Tel. #44 131 650 5345
FAX. #44 131 650 6563
Mobile 0410 585 401
email c.jeffree-at-ed.ac.uk
SEM / TEM bookings sem-at-ed.ac.uk
=====================================================================




From: A.P.Alves de Matos :      apmatos-at-ip.pt
Date: Sat, 3 Jun 2000 22:25:13 +0100
Subject: RE: TEM:beam damage vs voltage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I work with EPON sections of biological specimens. It seems clear to me that
60KV introduces more dammage than 80Kv. The sections are more unstable and
tend to break more easily at the lower voltage.

Dr. A.P. Alves de Matos
Pathology Department
Curry Cabral Hospital
Lisbon




------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Conventional wisdom said that when it came to using TEM for biological
specimens (or beam sensitive specimens) it is always best to use a lower
accelerating voltage. Many lower voltage microscopes (V { 120 kV) were sold
on this assumption.
However is this assumption true? At the present there are many research
establishments, buying TEMs for biological use, who are using 200 to 300 kV
beams. So obviously there has been a shift in the conventional way of
thinking. Being materials based I am not sure what the status quo is in
biological TEM.
I know that the ratio of inelastic to elastic scattering cross sections is
greater than one for the elements Z {12, but how does this change as the
beam energy increases? What are your experiences? This is really just
academic curiosity by the way, but I am sure many of you would appreciate
the question.
********************************************************
Dr Jonathan Barnard

Analytical Materials Physics
The Angstrom Laboratory, Uppsala University
P O Box 534, SE-751 21 Uppsala, Sweden
Phone: +46-(0)18-4716838 Fax: +46-(0)18-500131
Phone: Microscope room +46 18 471 6365
http://www.angstrom.uu.se/analytical/home.html
********************************************************








From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Sat, 03 Jun 2000 19:17:50 -0700
Subject: Re: Resolving power: film vs digital Commercial Post

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 02:21 PM 6/3/00, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I hope not.


} Digital light microscope cameras are now available that use a highly touted
} CCD detector: Color or monochrome; fast interline transfer, high sensitivity,
} low noise; 6.7 micron square pixel; 2/3 inch 1.3 Million chip; PCI bus frame
} transfer. This detector is ideal to use in light microscopy applications.
} Now precision technology and fast high capacity cheap computers allow this
} detector to be "Micro Stepped" providing digital images of up to 12 million
} pixels to be captured very quickly and providing file sizes of up to 35 MB in
} 24 bit color images. "Inter Pixel Stepping" technology will allow a
} considerably less expensive approach to digital imaging that can now approach
} film resolving capability in a light microscope.
}
} Hats of to Dr. Jeffree's explanation on optical resolution as it relates to
} light microscopy applications. With high pixel density capable digital
} cameras, now lower magnification, low NA (lower resolving power) objectives
} can be used to acquire matching digital resolution images of wide fields of
} view.
}
} Information on Nikon's Instrument Division "Digital Eclipse" family of
} digital cameras for microscopy including the new DXM1200 digital color camera
} will soon be available on our web site www.nikonusa.com or you can go to
} our Dealer Locator at
} http://www.nikonusa.com/corpinfo/dealers/dealerSearch.cfm and ask your
} local Nikon authorized microscope dealer for further information.
}
} Best Regards,
}
} Stan Schwartz
} Manager, BioSciences Dept.
} Nikon Inc Instrument Division
} 1300 Walt Whitman Rd.
} Melville, NY 11747
} 631-547-8500
} 631-547-4033 Fax
} Schwartz-at-nikonincmail.com
} www.nikonusa.com

As a long term past user of Nikon equipment, it is a tale of sorrow.
Most recently, the digital E1 and E2 are dismal failures. Consumers
are sending back 990's in droves (see rec.photo.marketplace.digital).
I really think that Nikon blew it in regards to digicams. How Nikon can
claim to get a 1.3M pixel imager to produce realistic 12M pixel images
is rather absurd....if not offensive. The most recent D1 is a joke. Albeit,
an expensive one.

Nikon blew it in the scanner arena and made huge blunders in high end
pro digicams (E1 & E2). Total junk from my personal experiences. I am
not at all prone to spend a dime on any new Nikon digital things. In fact,
I have
dumped my Nikon so-called pro lenses and bodies for Contax. But this
is another story.

My advice is to be very wary....be very wary of Nikon. I do not use Nikon
cameras anymore and I do not use Nikon microscopes anymore.
But it is your money and your decision. As they say, "Caveat emptor."

gg

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Modern surfers use PC boards. You can too at
http://photoweb.net
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: Chris Jeffree :      cjeffree-at-srv0.bio.ed.ac.uk
Date: Sun, 4 Jun 2000 09:49:36 +0100
Subject: Re: Resolving power: film vs digital Commercial Post

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rubbish! With all due respect this is a complete distortion and mis-
use of the point I made. To realise this objective with low NA lenses
you would have to find some way to defeat the laws of physics. No
CCD imager, however clever will make it possible to correct the
shortcomings of cheap, low-performance optics. You clearly
misunderstood the point, which is that microscope lens resoltuion is
not fundamentally dependent on the magnification factor but on the
numerical aperture, which also determines resolution. A x100 NA
1.4 lens resolves no more detail than a x60 NA 1.4, but simply
magnifies the image further. This is "empty magnification", which is
only useful if your image sensor (film or CCD) has limited resolving
power. With a very high resolution film like Technical Pan, you can
take advantage of this fact to capture a wide-field diffraction-limited
image with a x60 1.4 lens. That is not an option with a low NA lens
irrespective of the properties of the sensor, CCD or otherwise.

} Hats of to Dr. Jeffree's explanation on optical resolution as it relates to
} light microscopy applications. With high pixel density capable digital
} cameras, now lower magnification, low NA (lower resolving power) objectives
} can be used to acquire matching digital resolution images of wide fields of
} view.
}
} Information on Nikon's Instrument Division "Digital Eclipse" family of
} digital cameras for microscopy including the new DXM1200 digital color camera
} will soon be available on our web site www.nikonusa.com or you can go to
} our Dealer Locator at
} http://www.nikonusa.com/corpinfo/dealers/dealerSearch.cfm and ask your
} local Nikon authorized microscope dealer for further information.
}
} Best Regards,
}
} Stan Schwartz
} Manager, BioSciences Dept.
} Nikon Inc Instrument Division
} 1300 Walt Whitman Rd.
} Melville, NY 11747
} 631-547-8500
} 631-547-4033 Fax
} Schwartz-at-nikonincmail.com
} www.nikonusa.com
}
}
} Earlier post
} =============================================================
} Geoff
} Analysing this a little further:
} A 25 Mbyte image = 8.3 M pixels to cover 24x36mm or 864mm^2
} this represents 98.2 pixels per mm, or 49 line pairs per mm. Pixel
} size 10.2 um. Resolution about 50 line pairs per mm at best.
}
} A 25M pixels image used to capture a 24x36mm Kodachrome
} slide represents 28,935 pixels per mm^2 or 170 pixels per mm.
} This is equivalent to a maximum resolution of 85 line pairs per mm,
} which may be on the conservative side for Kodachrome. pixel size
} = 5.88um. The image will be approx. 6120x4082 pixels, generating
} a file size of approx 75Mb for a 24-bit (8+8+8bit) RGB image.
}
} To record 120 line pairs per mm, which many top 35mm camera
} lenses can achieve, a minimum of 240 pixels per mm are required,
} each 4.2 um wide. This equates to 8640x5760 pixels for a
} 24x36mm frame = 50M-pixels or 50Mb in 8-bit greyscale, or 150Mb
} in 24-bit RGB.
}
} At 320 line pairs per mm (Technical Pan) the minimum required
} 640 pixels per mm is a pixel size of 1.56 um
}
} Presumably for a light image the diffraction limited resolution is
} approx 1/2 lambda which at 540nm is 0.27um.
} So looking to the future of ultimate-performance CCDs, direct
} recording of a diffraction limited light image projected onto the
} sensor requires at the very least 3703 pixels per image mm or
} 13,717,421 pixels per mm^2 (greyscale 8-bit)
}
} However, if we are doing light microscopy with an NA 1.4 x100
} objective, how much resolving power do we need on CCD or film?
}
} Data is at 0.27um resolution (lambda = 540nm). Let's round this to
} 0.3um. Magnification at 24x36mm film image is x100, so pixels
} must be an absolute maximum of 30um wide to record the
} significant data = 33.3 pixels per mm, equivalent to 800x1200
} pixels to record the whole 35mm negative area. However, most
} CCDs are much smaller than 35 mm frames, typically 1/3 inch. So
} an 800x1200 pixel CCD at 8x12 mm, 1/3 of the linear dimensions
} of a 35mm frame would use 9 pixels to record the smallest image
} details. This is about right from the point of view of resolution, but
} to record the whole 35mm frame we need about 2400x3600 pixels
} on our CCD.
}
} Note also that Technical Pan has (depending on the criterion used
} to assess its performance) up to 10 times the resolving power
} required to record all there is to see in a diffraction-limited LM
} image made with a 100x NA 1.4 lens. So you can comfortably
} afford to use a 60x NA 1.4 lens, thereby getting the same
} resolution with a bigger field of view.
}
} Many years ago, I took a photograph of a street scene using a
} Canon 35mm SLR loaded with Kodak Recordak (I think this was a
} single layer microfilm emulsion). Examined in a light microscope,
} the image clearly, legibly recorded the brand-name of a child's
} push chair. I tried to print this brand name using a DeVere point-
} source enlarger with an image size of 20x30 inches produced with
} a Schneider Componon lens, but was completely unable to
} produce a legible image. The point I am making here is that the
} combination of some high performance films, with high quality
} lenses of the standard produced by the leading camera
} manufacturers can record more detail on the film than you can
} easily get back out by conventional printing. I suspect the same is
} true of EM exposures.
}
} So there is no contest - film beats CCDs for resolution hands
} down. And you can process the image on the cheap. No money
} goes to Intel or Microscoft, Adobe or Epson. But resolution is not
} primarily what we buy CCDs for. We buy them primarily for instant
} image capture in a format suitable for digital storage, digital
} transmission quantitative data recording and image processing.
}
} Chris
}
} =====================================================================
} DR CHRIS JEFFREE
} BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
} UNIVERSITY OF EDINBURGH
} Daniel Rutherford Building
} King's Buildings, Mayfield Road
} EDINBURGH, EH9 3JH, Scotland, UK
} Tel. #44 131 650 5345
} FAX. #44 131 650 6563
} Mobile 0410 585 401
} email c.jeffree-at-ed.ac.uk
} SEM / TEM bookings sem-at-ed.ac.uk
} =====================================================================
}


~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Dr Chris Jeffree
University of Edinburgh
Biological Sciences EM Facility
Daniel Rutherford Building
King's Buildings EDINBURGH EH9 3JH
Tel: +44 (0) 131 650 5345
FAX: +44 (0) 131 650 6563

Inveresk Cottage, 26 Carberry Road,
Inveresk, Musselburgh, Midlothian EH21 8PR, UK
Tel. +44 (0) 131 665 6062 / Mobile 0410 585 401
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~




From: jim :      jim-at-proscitech.com.au
Date: Sun, 4 Jun 2000 21:20:44 +1000
Subject: RE: Film vs Digital/ switched to EM techn

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lets place the exhausted and exhausting Digital topic aside.
Sergey, you do have a challenging project and one that is worth discussing.
Just maybe somebody has an idea that will help you. I think that we discuss in
this forum pixels too much and microscopy too little.

You probably have tried most of my following suggestions, but just incase, here
are a couple of my thoughts:

Dark field contrast is enhanced most when the density between specimen and
background is greatest. So its most effective with no specimen support. You
could try your luck with the superfine mesh grids now available; these thin bar
grids can be purchased down to 2000 mesh and these grids have a 7.5um hole
size. If that is not a fine enough support, then try holey films with a
net-like structure.

Carbon coating will stabilize samples dramatically. A little loss of contrast
is inevitable, but a least carbon on the grid or holey plastic film does not
matter.

Don't see why you want to render the gold nano particles with detail (I assume
some greys). In TEM I would expect gold above about 5nm to be black. In STEM or
FESEM you could get greys easily, but you may not get the resolution required.

Increasing if available to 200 or more the kV, will give more brightness, less
contrast, better resolution (specimen dependent). Most importantly, specimen
damage frequently is less at higher kV, that depends on the specimen again. I
think that damage is less in specimens with greater electron transparency, but
one of our "physical" gurus may explain the whys and wherefores.

Without switching film-types, you can develop the TEM films in things like
Microdol-X, Microphen or Rodinal (hope those developers still exist). You could
also use D-19 more dilute than normal. For most EM purposes all of these would
give too softer negatives, but they just may suit you. Over developing in light
photography yields more contrast - not so in TEM, where more electrons are the
main contrast mechanism. Hence slower film and denser negatives are preferred,
especially by most biologists.

I think that we started out with too much contrast, but a grainy image because
of insufficient electron exposure. Hmmm, I think that we should take a holiday.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Saturday, June 03, 2000 7:44 AM, Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
wrote:
}
}
} Dear Jim
}
} I spent about two years trying to make good pictures of my NanoGold labeled
} protein-DNA complexes. Doing this job I find two main problems: the
} sample is unstable under the beam as any biological sample; NanoGold is
} much more bright than protein core in dark-field. I find that it is
} impossible to record equally perfect signals from NanoGold and protein core
} because of short dynamic range for SO-163 film, I believe. I was trying to
} make two pictures with different exposure, but it is tricky: in dark-field
} mode the automatic exposure meter usually does not work and we have to set
} exposure manually, in this case it is difficult to get "right" exposure
} time in the right moment, you know. Again, because of sample's short life
} under the beam, it is impossible to make a couple pictures at the different
} conditions sometime. Keep in mind, please, that to change the film in the
} microscope it takes about 10 seconds. Your idea about increasing
} signal-noise ratio by collecting more electrons is bright but not
} practical. For biological samples (I am talking about non-fixed,
} non-stained samples of proteins, DNA or RNA-protein complexes etc) the
} electron damage is a huge problem. People are trying to solve it in
} different ways. Some using cryo temperature (to stabilize the biological
} structure). I was using freeze-drying (I find that freeze-dried samples
} are more stable under the beam). But in any case we have deal with very
} unstable samples and must to do everything to decrease (not increase as you
} recommended) electron dose. Drift is a second big problem for such
} application: to increase signal-noise ratio we have to use very thin
} support films. Images obtained at such conditions are noisy and in most
} cases we have to use image analysis tools to extract the data. It means
} that we have to digitize our images anyway. In such situation digital
} camera may help. As you, probably, remember I was a person who initiates
} this discussion. I think this discussion was very useful for many of us
} who are not friendly with digital camera's techniques. We understand the
} limitations of the modern digital cameras better now. I would like to say
} thank you everybody who was involved in this discussion. There is some
} conclusions I make for myself from discussion:
}
}
} - Film is still cheap and universal material for recording and
} storage EM images, sorry CCD.
} - CCD TEM camera should not substitute film. Film and camera should
} work all together improving the flexibility of the TEM system. For this
} reason I will chose side-mount camera if will have money for it.
} - For cell-biology (thin sections) where the resolution of the sample
} is about 3 nm CCD camera may do a good job allowing users to make a huge
} number of pictures (cell-biology guys love it), instantly view and
} catalogize them.
} - Sometime the digital camera may help in area of high-resolution
} (relatively high, guys) EM when image will be digitized anyway. The major
} limitation here is small area of view (we need a lot of particles for image
} analysis sometime), but you could make the set of overlapping pictures and
} digitally combine it. I love, also, Mike Bode idea to make a few very
} short exposure pictures and combine it digitally later to reduce noise. The
} relatively big size of CCD's pixels is a real problem too.
} - CCD camera is expensive "toy". I am not sure that the benefits from
} using it will compensate astronomical price, actually the third of the
} electron microscope value ($70000 is it 1/3 of microscope's price on
} current market?). Currently, I would recognize the CCD TEM camera as funny
} "attachment" which may be useful if you rich enough to spend money on it
} (it mean, that you have everything else in your EM lab plus some extra
} $70000 for the fun playing with digital "toy").
} - TEM CCD cameras are under extensive development now. Today's
} camera will be replaced on the new model (read better, faster, what else?)
} next year. Next year's camera will be easily forgotten next after the next
} year and so on... Each new camera will be better that previous one... CCD
} TEM camera it is not a good investment of money, I think.
} - We should keep in mind that many companies charged extra 3-4K$ for
} the installation and training (it is mandatory for Gatan for instance) and
} you, probably, have to buy service contract on it even if you have service
} contract on the microscope (JEOL's service contract on microscope do not
} cover the CCD camera even if you buy it from JEOL).
}
} Best regard, Sergey
}
}
}
}
} } Date: Fri, 02 Jun 2000 21:14:44 +1000
} } From: jim {jim-at-proscitech.com.au}
} } Subject: RE: Film vs Digital
} } To: 'Mike Bode' {mb-at-Soft-Imaging.com} ,
} } "'Microscopy-at-MSA.Microscopy.Com'" {Microscopy-at-sparc5.microscopy.com}
} } Cc: "'jim-at-proscitech.com.au'" {jim-at-proscitech.com.au}
} } Reply-to: "jim-at-proscitech.com.au" {jim-at-proscitech.com.au}
} } Organization: ProSciTech
} } X-Mailer: Microsoft Internet E-mail/MAPI - 8.0.0.4211
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } The world is full of possible solutions, but are they practical.?
} } To produce high-resolution, dark-field or any others TEM images that require
} } more electrons to form a clear image, Mike Bode would use multiple digital
} } exposures. The exposures could be layered and combined into one superior
} } image.
} } This image would be made up of more pixel and is formed by more electrons
} } and
} } so would be noise-free and hence could be further enlarged then otherwise
} } possible. Perhaps.
} } Beam blanking would largely save the specimen from beam damage and drift
} } could
} } be compensated for by matching up the digitals. Great.
} } How much time is required between exposures to transfer a minimum 10mb image
} } per exposure? What would be the total time from focusing to the last
} } exposure?
} } What about Z-drift in the interim requiring objective changes and what about
} } the total cost of this additional get-up. The mind boggles at a through
} } focus
} } series.
} } When pushing the limits a piece of film seems more effective, cheaper and fa
} } ster.
} } Again, I don't doubt that there is now a large place for digital in TEM, but
} } its no panacea.
} } Cheers
} } Jim Darley
} } ProSciTech Microscopy PLUS
} } PO Box 111, Thuringowa QLD 4817 Australia
} } Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} } Great microscopy catalogue, 500 Links, MSDS, User Notes
} } www.proscitech.com
} }
} } On Friday, June 02, 2000 1:24 AM, Mike Bode [SMTP:mb-at-soft-imaging.com]
wrote:
} }
} } }
} } } No, it's not really a problem. It's been done with low density
} } } microscopy all the time. Granted, there are some technical aspects to be
} } } overcome, but (and I can only speak for ourselves) we have done that on
} } } a number of microscopes. You are of course correct, that 10 images at
} } } 0.8 seconds take longer than 8 seconds as the image has to be
} } } transferred, etc. BUT: that's what beam blankers are for. It is pretty
} } } straightforward to take an image at 0.8 seconds, then blank the beam
} } } very quickly before taking the next image. That way you get pretty close
} } } to the 8 sec total exposure. If there is no beam blanker on the
} } } microscope, in most cases it can be added.
} } }
} } } I am not sure what you mean by "too sensitive". The cameras are usually
} } } constructed so that 1 electron from the beam creates between a few tenth
} } } to a few counts (these are all statistical data, of course). The well
} } } width divided by this sensitivity then determines, how many primary
} } } electrons are needed to fully expose one pixel. For example, if the well
} } } width is 50,000 electrons and the sensitivity is 1 count/electron, one
} } } needs 50,000 primary electrons to fill the well. This translates into
} } } roughly a 0.4% statistical error.
} } }
} } } } From a practical standpoint: You can take images with most cameras when
} } } the exposure meter on the microscope reads a couple of seconds without
} } } overexposing the camera. On the other hand, you can reduce the intensity
} } } of the beam until you see single electron events.
} } }
} } } The one area where CCD cameras may be too sensitive is diffraction. The
} } } normally huge intensity in the transmitted beam often leads to
} } } saturation. In CCDs this can lead to blooming (the intensity spills over
} } } into neighboring pixels). This can be taken care of with special chips
} } } that have anti-blooming features, but this usually has some other
} } } drawbacks. Again, this can also be overcome somewhat with multiple
} } } exposures. Film behaves more civilized here, as it simply stops
} } } responding to the electrons, but this makes film more or less useless
} } } for quantitative measurements of diffraction patterns. I have done
} } } diffraction with CCDs many times and though it does require some
} } } tweaking, one can get very good results from them.
} } }
} } } Michael Bode, Ph.D.
} } } Soft Imaging System Corp.
} } } 1675 Carr St., #105N
} } } Lakewood, CO 80215
} } } ===================================
} } } phone: (888) FIND SIS
} } } (303) 234-9270
} } } fax: (303) 234-9271
} } } email: mailto:info-at-soft-imaging.com
} } } web: http://www.soft-imaging.com
} } } ===================================
} } }
} } }
} } }
} } } -----Original Message-----
} } } From: jim [mailto:jim-at-proscitech.com.au]
} } } Sent: Wednesday, May 31, 2000 9:37 PM
} } } To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} } } Subject: RE: Film vs Digital
} } }
} } }
} } } ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------.
} } }
} } }
} } } Yes, it could be done "in theory". Somebody would need to figure out the
} } }
} } } software and perhaps modify the hardware. Then we would find that the
} } } total
} } } exposure of the specimen to the electron beam maybe a muliple of the
} } } film's
} } } exposure. Afterall, an 8 sec film exposure would not amount in digital
} } } to
} } } 10x0.8, but we would require considerable time in between exposures.
} } } Since the
} } } problems in the discussed circumstances are specimen movement and beam
} } } damage,
} } } it seems that taking multiple exposures is a poor option.
} } }
} } } Digital cameras are for some situation too sensitive to electron
} } } exposure.
} } } Cutting back on electrons is no option since its the electrons that form
} } } the
} } } image in the first instance.
} } } Much easier in light microscopy . . . insert a neutral density filter.
} } } Cheers
} } } Jim Darley
} } } ProSciTech Microscopy PLUS
} } } PO Box 111, Thuringowa QLD 4817 Australia
} } } Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} } } Great microscopy catalogue, 500 Links, MSDS, User Notes
} } } www.proscitech.com
} } }
}
} _____________________________________
}
} Sergey Ryazantsev Ph. D.
} Electron Microscopy
} UCLA School of Medicine
} Department of Biological Chemistry
} Box 951737
} Los Angeles, CA 90095-1737
}
} Phone: (310) 825-1144
} Pager: (310) 845-0248
} FAX (departmental): (310) 206-5272
} mailto:sryazant-at-ucla.edu
} http://www.bol.ucla.edu/~sryazant
}
}





From: jim :      jim-at-proscitech.com.au
Date: Sun, 4 Jun 2000 18:33:42 +1000
Subject: RE: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have two replies and Mike my well have four replies to these.
1. There has been a parallel discussion concerning resolution of film versus
digital images. Clearly in raw power digital cannot compete since film has
multi gigabyte capacity. I added to that thread that what matters is: does
digital have enough power and that frequently it would. Mike reinforced and
strengthened that argument, finishing with the note that he is glad for film
when looking at prints by Ansel Adams.
(Ansel Adams until about 30 years ago carted for decades large format cameras
through US National Parks, especially Yosemite, producing fantastic landscape
photographs and books)
Adams' limited edition prints were contacts of 5x7 and 8x10" inch sheet film,
probably rated at 400 ISO. The line resolution of such prints much exceeds our
eyes' resolution, but still results in superior gradation and detail. TEM film,
even when much enlarged has such details too.
Why Mike, should we accept 2.5 Mb? That is a splendid file size for SEM, and
because of the limited enlargability of light microscopy (concrete ceiling due
to wavelengths) it's reasonable, but minimal for light microscopy. TEM can do
and deserves better.

Incidentally, from the outset I cited increased "enlargability" to obtain high
resolution TEM, as one of films major advantages, since greater depths of field
at moderate powers makes high powers through photo enlarging a desirable
technique. The small additional magnification yielded by placing a digital
camera lower in the column does not compensate. So greater "enlargability" of
digitals would be desirable, but is limited by pixel size.

2 The thread was initiated by Sergey. He had problems visualising certain
specimens in dark field. The use of his "beaut" digital TEM camera made things
worse. I pointed out that the shorter exposure reduced the number of electrons
forming the image, hence more noise. I believe that a good part of Mike's case
will be settled in his favour when we hear "Eureka" from Sergey's lab. I don't
doubt that much more can be done with digital now and that further improvements
are on the way. Mike's "solution" may well be possible, but I don't believe its
a snap; in any case: Show Sergey!
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Saturday, June 03, 2000 2:58 AM, Mike Bode [SMTP:mb-at-Soft-Imaging.com] wrote:
}
} Well, let's see:
}
} Jim wrote:
}
} Mike Bode would use multiple digital
} exposures. The exposures could be layered and combined into one superior
} image.
} This image would be made up of more pixel and is formed by more
} electrons and
} so would be noise-free and hence could be further enlarged then
} otherwise
} possible. Perhaps.
}
} No, I did not talk about further enlargements. All I wanted to say is,
} that a more noise-free image can be achieved by adding multiple images,
} and that this also to some extent helps with drift of the sample during
} acquisition.
}
} Jim wrote:
}
} How much time is required between exposures to transfer a minimum 10mb
} image
} per exposure?
}
} How did you arrive at 10 MB? A 1280x1024 image with 16 bit pixel
} information is about 2.5 MB (uncompressed). We acquire about 10 of those
} per second and transfer them across the PC bus to the display. Putting
} them on them into Memory might add a few tenth of a second. Writing to
} HD can be done after all images are acquired.
}
} Jim wrote:
}
} What would be the total time from focusing to the last exposure?
} What about Z-drift in the interim requiring objective changes
}
} Why would we have to worry about that, if we don't have to worry about
} that when taking the image on film? In fact, we could take care of this
} by looking at the image between exposures and correct for z-drift.
} However, as you said, that would add to the overall time and exposure. I
} was comparing a normal dark field image taken on film at 8 seconds with
} acquiring the same image on a "too sensitive" CCD camera by adding up 10
} consecutive .8 second images. Why would the sample drift (in x, y or z)
} substantially more in 8+delta seconds than in 8?
}
} Jim wrote:
}
} what about the total cost of this additional get-up
}
} That of course depends on the microscope and there is no general answer.
} For example on a LEO 912 I believe the blanker is standard. The
} additional cost to use an acquisition scheme like this with our software
} is $0 plus perhaps a bit of time to write a small macro. On other
} microscopes one might have to add a beam blanker and perhaps a control
} mechanism for the beam blanker. But I would guess, that this cost is not
} very high. All modern microscopes are computer controller anyway, so it
} is most likely just a control command that needs to be sent to the
} microscope over a serial port if the beam blanker is installed. Piece of
} cake.
}
} Jim wrote:
}
} The mind boggles at a through focus series.
}
} You're right here. But I don't think we were talking about through-focus
} series. Incidentally, we do through-focus series on light microscopes
} and reconstruction routinely. Takes a few images at different focus (or
} for a light microscope: stage) settings. The rest is done off-line.
} Takes maybe a couple of minutes for about 20 images of about 1kx1k. I
} agree that TEM is different here and much more complicated due to the
} complicated Contrast Transfer Function. However, this could in principle
} be sorted out.
}
} Jim wrote:
}
} Again, I don't doubt that there is now a large place for digital in TEM,
} but
} its no panacea.
}
} I also agree with you on that one. But using the additional computer
} possibilities of digital imaging might take you further than expected.
}
} Michael
}
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 1675 Carr St., #105N
} Lakewood, CO 80215
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} } From: jim [mailto:jim-at-proscitech.com.au]
} Sent: Friday, June 02, 2000 5:15 AM
} To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} Cc: 'jim-at-proscitech.com.au'
} Subject: RE: Film vs Digital
}
}
} The world is full of possible solutions, but are they practical.?
} To produce high-resolution, dark-field or any others TEM images that
} require
} more electrons to form a clear image, Mike Bode would use multiple
} digital
} exposures. The exposures could be layered and combined into one superior
} image.
} This image would be made up of more pixel and is formed by more
} electrons and
} so would be noise-free and hence could be further enlarged then
} otherwise
} possible. Perhaps.
} Beam blanking would largely save the specimen from beam damage and drift
} could
} be compensated for by matching up the digitals. Great.
} How much time is required between exposures to transfer a minimum 10mb
} image
} per exposure? What would be the total time from focusing to the last
} exposure?
} What about Z-drift in the interim requiring objective changes and what
} about
} the total cost of this additional get-up. The mind boggles at a through
} focus
} series.
} When pushing the limits a piece of film seems more effective, cheaper
} and fa
} ster.
} Again, I don't doubt that there is now a large place for digital in TEM,
} but
} its no panacea.
} Cheers
} Jim Darley
} ProSciTech Microscopy PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} www.proscitech.com
}
} On Friday, June 02, 2000 1:24 AM, Mike Bode [SMTP:mb-at-soft-imaging.com]
} wrote:
} }
} } No, it's not really a problem. It's been done with low density
} } microscopy all the time. Granted, there are some technical aspects to
} be
} } overcome, but (and I can only speak for ourselves) we have done that
} on
} } a number of microscopes. You are of course correct, that 10 images at
} } 0.8 seconds take longer than 8 seconds as the image has to be
} } transferred, etc. BUT: that's what beam blankers are for. It is pretty
} } straightforward to take an image at 0.8 seconds, then blank the beam
} } very quickly before taking the next image. That way you get pretty
} close
} } to the 8 sec total exposure. If there is no beam blanker on the
} } microscope, in most cases it can be added.
} }
} } I am not sure what you mean by "too sensitive". The cameras are
} usually
} } constructed so that 1 electron from the beam creates between a few
} tenth
} } to a few counts (these are all statistical data, of course). The well
} } width divided by this sensitivity then determines, how many primary
} } electrons are needed to fully expose one pixel. For example, if the
} well
} } width is 50,000 electrons and the sensitivity is 1 count/electron, one
} } needs 50,000 primary electrons to fill the well. This translates into
} } roughly a 0.4% statistical error.
} }
} } } From a practical standpoint: You can take images with most cameras
} when
} } the exposure meter on the microscope reads a couple of seconds without
} } overexposing the camera. On the other hand, you can reduce the
} intensity
} } of the beam until you see single electron events.
} }
} } The one area where CCD cameras may be too sensitive is diffraction.
} The
} } normally huge intensity in the transmitted beam often leads to
} } saturation. In CCDs this can lead to blooming (the intensity spills
} over
} } into neighboring pixels). This can be taken care of with special chips
} } that have anti-blooming features, but this usually has some other
} } drawbacks. Again, this can also be overcome somewhat with multiple
} } exposures. Film behaves more civilized here, as it simply stops
} } responding to the electrons, but this makes film more or less useless
} } for quantitative measurements of diffraction patterns. I have done
} } diffraction with CCDs many times and though it does require some
} } tweaking, one can get very good results from them.
} }
} } Michael Bode, Ph.D.
} } Soft Imaging System Corp.
} } 1675 Carr St., #105N
} } Lakewood, CO 80215
} } ===================================
} } phone: (888) FIND SIS
} } (303) 234-9270
} } fax: (303) 234-9271
} } email: mailto:info-at-soft-imaging.com
} } web: http://www.soft-imaging.com
} } ===================================
} }
} }
} }
} } -----Original Message-----
} } From: jim [mailto:jim-at-proscitech.com.au]
} } Sent: Wednesday, May 31, 2000 9:37 PM
} } To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} } Subject: RE: Film vs Digital
} }
} }
} }
} ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} -----------------------------------------------------------------------.
} }
} }
} } Yes, it could be done "in theory". Somebody would need to figure out
} the
} }
} } software and perhaps modify the hardware. Then we would find that the
} } total
} } exposure of the specimen to the electron beam maybe a muliple of the
} } film's
} } exposure. Afterall, an 8 sec film exposure would not amount in digital
} } to
} } 10x0.8, but we would require considerable time in between exposures.
} } Since the
} } problems in the discussed circumstances are specimen movement and beam
} } damage,
} } it seems that taking multiple exposures is a poor option.
} }
} } Digital cameras are for some situation too sensitive to electron
} } exposure.
} } Cutting back on electrons is no option since its the electrons that
} form
} } the
} } image in the first instance.
} } Much easier in light microscopy . . . insert a neutral density
} filter.
} } Cheers
} } Jim Darley
} } ProSciTech Microscopy PLUS
} } PO Box 111, Thuringowa QLD 4817 Australia
} } Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} } Great microscopy catalogue, 500 Links, MSDS, User Notes
} } www.proscitech.com
} }





From: Lawrence Kordon :      nikon-at-jagunet.com
Date: Sun, 04 Jun 2000 09:53:36 -0400
Subject: Re: Resolving power: film vs digital Commercial Post

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


****This is a commercial response from a Vendor****

Gary and List,

I am sorry you have had a bad experience with Nikon products. However, you are
entitled to your "opinion" and therefore; I am entitled to a response. First, the
new Digital Eclipse Camera capable of 12M pixels using a stepper mode is exactly
the same technology that the new Zeiss Axiocam and an Olympus model uses which has
been very well received by the microscope community.

Next, the successor to E1, E2 (technically, made by Fuji) is the current Nikon D1
(Digital SLR) was NOT intended for microscopes. It CAN go on one, but is limited
to Brightfield applications and has no NTSC (video) out for focusing. However, it
is by far the standard in Digital Cameras! It has won every Journal Award in it's
field and is the definitive choice for Professional Photo-Journalists, including
the massive market share Nikon has and the majority of Pulitzer Prize winners. As
for Film Scanner products, the Nikon Coolscan line is still the standard in the
industry. Especially, high end units with Auto Feeders. All Nikon digital products
are not perfect. However, these are technical products that are evolving 6 months
at a time and are truly in their infantile state from where they will be in just a
few years from now! I would prefer to address your specific problems, but you gave
none and just called it "junk". With that mentality, I feel my Pentium 266
computer is "junk", but I do not blame the manufacturer because I own an older
model.

As for the Nikon Coolpix line (990), pardon my sarcasm, but where are "all the
returns" because we need them to fulfill the 38,000 unit Backorders! This camera
is so wildly successful EVERY Dealer (Photo and Microscope) is begging for them.
Check out Ebay for example; the cameras are selling for more than List Price.
Sound to me like a success if you understand the basics of Economics and Supply
and Demand. Is it the perfect microscope camera? No, I don't even think so...but
it is the ONLY high resolution (3.34M) digital camera, with live video out, goes
on any brand or model microscope for under $1000. Oh, did I forget to mention it
also won almost every Magazine Award in its class (not all, it actually received a
tie with the Olympus 3030 in one, which is a very nice camera, but does not mount
a microscope).

Finally, I have been on this List for many years and have respected the use and
rules of this forum. We Vendors for the most part are respectful of the opinions
of our users. However, making a "blanket statement" like "stay away from Nikon",
does not serve the public well; NOR YOURSELF! If you or any customer has specific
problems; we want to know about it; as does any reputable manufacturer for future
product improvement. The fact is Nikon has the largest market share in microscopes
and professional camera equipment and is climbing very quickly on the digital
camera list too. So, in short you are entitled to your opinion, but the rest of
the world by far does NOT agree with it............

I personally apologize to any members this correspondence offends. Believe me, my
preference is to serve the list and answer technical microscopy questions as I
have for many years, but this unfair and libel attack required a response.

Regretfully,

Lawrence Kordon
Nikon, Inc.
Senior Bioscience Specialist
nikon-at-jagunet.com

"Dr. Gary Gaugler" wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} At 02:21 PM 6/3/00, you wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Hello all,
} }
} } Please be aware that this is a commercial post that may be of interest to
} } some of the list members.
}
} I hope not.
}
} } Digital light microscope cameras are now available that use a highly touted
} } CCD detector: Color or monochrome; fast interline transfer, high sensitivity,
} } low noise; 6.7 micron square pixel; 2/3 inch 1.3 Million chip; PCI bus frame
} } transfer. This detector is ideal to use in light microscopy applications.
} } Now precision technology and fast high capacity cheap computers allow this
} } detector to be "Micro Stepped" providing digital images of up to 12 million
} } pixels to be captured very quickly and providing file sizes of up to 35 MB in
} } 24 bit color images. "Inter Pixel Stepping" technology will allow a
} } considerably less expensive approach to digital imaging that can now approach
} } film resolving capability in a light microscope.
} }
} } Hats of to Dr. Jeffree's explanation on optical resolution as it relates to
} } light microscopy applications. With high pixel density capable digital
} } cameras, now lower magnification, low NA (lower resolving power) objectives
} } can be used to acquire matching digital resolution images of wide fields of
} } view.
} }
} } Information on Nikon's Instrument Division "Digital Eclipse" family of
} } digital cameras for microscopy including the new DXM1200 digital color camera
} } will soon be available on our web site www.nikonusa.com or you can go to
} } our Dealer Locator at
} } http://www.nikonusa.com/corpinfo/dealers/dealerSearch.cfm and ask your
} } local Nikon authorized microscope dealer for further information.
} }
} } Best Regards,
} }
} } Stan Schwartz
} } Manager, BioSciences Dept.
} } Nikon Inc Instrument Division
} } 1300 Walt Whitman Rd.
} } Melville, NY 11747
} } 631-547-8500
} } 631-547-4033 Fax
} } Schwartz-at-nikonincmail.com
} } www.nikonusa.com
}
} As a long term past user of Nikon equipment, it is a tale of sorrow.
} Most recently, the digital E1 and E2 are dismal failures. Consumers
} are sending back 990's in droves (see rec.photo.marketplace.digital).
} I really think that Nikon blew it in regards to digicams. How Nikon can
} claim to get a 1.3M pixel imager to produce realistic 12M pixel images
} is rather absurd....if not offensive. The most recent D1 is a joke. Albeit,
} an expensive one.
}
} Nikon blew it in the scanner arena and made huge blunders in high end
} pro digicams (E1 & E2). Total junk from my personal experiences. I am
} not at all prone to spend a dime on any new Nikon digital things. In fact,
} I have
} dumped my Nikon so-called pro lenses and bodies for Contax. But this
} is another story.
}
} My advice is to be very wary....be very wary of Nikon. I do not use Nikon
} cameras anymore and I do not use Nikon microscopes anymore.
} But it is your money and your decision. As they say, "Caveat emptor."
}
} gg
}
} ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
} Modern surfers use PC boards. You can too at
} http://photoweb.net
} ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: Nestor J. Zaluzec :      zaluzec-at-sparc5.microscopy.com
Date: Sun, 4 Jun 2000 12:26:35 -0500
Subject: Administrivia: Resolving power: film vs digital Commercial Post

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues.....

Stop this "tangental thread" on Resolving Power now, before it
gets out of hand.
User/Vendor problems should be handled in private not on the list.


Nestor
Your Friendly Neighborhood SysOp.






From: Caroline Schooley :      schooley-at-mcn.org
Date: Sun, 4 Jun 2000 16:29:41 -0700
Subject: Project MICRO

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


It's been some time since I posted news about Project MICRO, MSA's middle
school educational outreach program. I'm happy to report that Nestor,
wearing his Webmaster hat (yes, he owns more than one), has just posted a
substantial revision of the MICRO website (URL below). You'll find new
information on several pages and a LOT of new entries in the bibliography.
Don't miss the "Cyclops" videos and the new CD-ROMs; the website hotlinks
are much expanded also.

Caroline

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: John E. Toth :      jetoth-at-mail.olypen.com
Date: Sun, 4 Jun 2000 21:53:51 -0500
Subject: Microsope design

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sirs,

I am looking for books on the subject of optical microscope objective
lens design.
Any suggestions would be appreciated.

John Toth

jetoth-at-olypen.com
or
sciret-at-olypen.com






From: Toby Knight :      tknight-at-waite.adelaide.edu.au
Date: Mon, 5 Jun 2000 12:29:02 +0930 (CST)
Subject: schiffs staining epoxy sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I want to stain Procure-Araldite embedded material using the periodic-acid
schiff procedure. One method I have is to hydrolyze in 1N HCl at
60 degrees C for 10 mins, water wash, stain in schiffs, metabisulphate
rinses, counterstain, water wash (O'Brien and McCully, 1981). Does anyone
use a method routinely which varies from this??

thanks in advance, Toby Knight.


--------------------------------------------------------
Toby Knight
PhD student

Department of Horticulture, Viticulture and Oenology
The University of Adelaide
Plant Research Centre
Waite Campus, PMB 1
Glen Osmond, SA 5064

Tel: +61 8 8303 7224 or 8303 6668
HVO: +61 8 8303 7242
Fax: +61 8 8303 7116
Email: tknight-at-waite.adelaide.edu.au
--------------------------------------------------------







From: MGMANDERS-at-aol.com
Date: Sun, 4 Jun 2000 22:41:14 -0500
Subject: NEED HELP WITH EDS ON A JEOL 100S

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A fellow colleague has bought a JEOL 100s and wants and needs EDS X-ray
Analysis. He's mounted his horizintal EDS detector , but can not get sample
X-rays from the speciman. Contacting JEOL he found the problem to be a tilt
problem. The 10 degree tilt from the normal SEG only tilts 10 degrees and a
30 to 60 degree tilt is necessary. At one time I was told that special
sample holders and or double gap pole pieces were availabe. He wishs to find
any one with a 100s for parts needed or information which would allow X-ray
analysis. Please reply to mgmanders-at-aol.com or the list server.

Mike Manders






From: Gordon Couger :      gcouger-at-couger.com
Date: Sun, 4 Jun 2000 23:46:40 -0500
Subject: Desktop incubator

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I reacall some one looking for a desktop incubator. I came
a across one on ebay.
http://cgi.ebay.com/aw-cgi/eBayISAPI.dll?ViewItem&item=344913999
It is a little pricy by ebay standards. You can find a bunch of incubators
by going to www.ebay.com and search for incubators you can do the
same a http://www.labx.com

I don't have any connection with anyone involved in this.

Gordon

Gordon Couger gcouger-at-couger.com

Stillwater, OK www.couger.com/gcouger
405 624-2855 GMT -6:00






From: John Nailon :      mmjnailo-at-dingo.cc.uq.edu.au
Date: Sat, 03 Jun 2000 16:35:28 +1000
Subject: Re: schiffs staining epoxy sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Toby,
I used this technique years ago when I worked for Terry O'Brien at
Monash Uni. Depending on the type of tissue you may want to do an
aldehyde blockade before you begin the staining procedure. This
technique is also in O'Brien and McCully. The blockade removes any
aldehyde grouping that will react with the Schiffs reagent. You then
generate and stain aldehyde groups during the procedure.
Regards
JVN

Toby Knight wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} I want to stain Procure-Araldite embedded material using the periodic-acid
} schiff procedure. One method I have is to hydrolyze in 1N HCl at
} 60 degrees C for 10 mins, water wash, stain in schiffs, metabisulphate
} rinses, counterstain, water wash (O'Brien and McCully, 1981). Does anyone
} use a method routinely which varies from this??
}
} thanks in advance, Toby Knight.
}
} --------------------------------------------------------
} Toby Knight
} PhD student
}
} Department of Horticulture, Viticulture and Oenology
} The University of Adelaide
} Plant Research Centre
} Waite Campus, PMB 1
} Glen Osmond, SA 5064
}
} Tel: +61 8 8303 7224 or 8303 6668
} HVO: +61 8 8303 7242
} Fax: +61 8 8303 7116
} Email: tknight-at-waite.adelaide.edu.au
} --------------------------------------------------------

--
****************************************************
John V Nailon
Operations Manager
Centre for Microscopy and Microanalysis
The University of Queensland
St. Lucia Queensland 4072
Phone: +61-7-3365-4214
Fax: +61-7-3365-4422
WWW: http://www.uq.edu.au/nanoworld/allstaff.html#Nailon
****************************************************




From: Michelle.Taurino-at-aventis.com
Date: Mon, 5 Jun 2000 06:54:20 -0500
Subject: schiffs staining epoxy sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Toby-
I've worked with the PAS stain with methyl methacrylate embedded tissue cut
at 2 microns.
I used a kit from Polyscientific (NY, USA) which supplied me with all of the
necessary reagents (minus the ethanol and xylene). The protocol that was
supplied with the kit is for paraffin embedded material but I made
modifications to the protocol to accommodate methyl methacrylate.
If you are interested in the modified protocol, please contact me.
I should also mention that I automated this stain to save time and to
maximize quality.

Good luck!
Michelle Taurino
Aventis Pharmaceuticals
Bioimaging and Molecular Histology
Tel-908-231-3357
Fax-908-231-3962
e-mail: Michelle.Taurino-at-aventis.com


-----Original Message-----
} From: Toby Knight [mailto:tknight-at-waite.adelaide.edu.au]
Sent: Sunday, June 04, 2000 10:59 PM
To: Histonet; Microscopy list


I want to stain Procure-Araldite embedded material using the periodic-acid
schiff procedure. One method I have is to hydrolyze in 1N HCl at
60 degrees C for 10 mins, water wash, stain in schiffs, metabisulphate
rinses, counterstain, water wash (O'Brien and McCully, 1981). Does anyone
use a method routinely which varies from this??

thanks in advance, Toby Knight.


--------------------------------------------------------
Toby Knight
PhD student

Department of Horticulture, Viticulture and Oenology
The University of Adelaide
Plant Research Centre
Waite Campus, PMB 1
Glen Osmond, SA 5064

Tel: +61 8 8303 7224 or 8303 6668
HVO: +61 8 8303 7242
Fax: +61 8 8303 7116
Email: tknight-at-waite.adelaide.edu.au
--------------------------------------------------------






From: Dr Malcolm Roberts :      malc-at-rock.ru.ac.za
Date: Mon, 05 Jun 2000 14:01:28 +0200
Subject: BSE-Thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Just like to thank all those who responded to my squeal for help on BS
detector resolution. Would like to add that the South African microscopy
community is looking into establishing QA procedures and some of the
suggestions maybe very helpful in this regard.
Thanks,
Malc.

--
Dr MP Roberts Phone: [+27](0)46 603 8313
Dept of Geology Fax: [+27](0)46 622 9715
Rhodes University Cell: 083 4060 262 (usually off)
6140 Grahamstown e-mail: malc-at-rock.ru.ac.za
SOUTH AFRICA






From: De Pauw Bart :      bart.depauw-at-rug.ac.be
Date: Mon, 05 Jun 2000 15:21:00 +0200
Subject: SEM - Mycoplasms

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Listers,

I've been trying to look at Mycoplasms with SEM. Untill now, the results
are not what we hoped for. We've used polycarbonate filters to prepare the
Mycoplasms. Another way was preparing the Mycoplasms on agar. We have done
this already for enterococcus and we know that this technique works well.
But with the Mycoplasms we see almost nothing. It's like the Mycoplasms are
IN the Agar.
Has anyone have experience with MYCOPLASMS and SEM ???


De Pauw Bart
Ghent University
Faculty of Veterinary Medicine
Morphology
Salisburylaan 133
9820 Merelbeke
Belgium
Phone : 0032(0)9 264.77.19
Fax : 0032(0)9 264.77.90





From: Chris Jeffree :      cjeffree-at-srv0.bio.ed.ac.uk
Date: Mon, 5 Jun 2000 16:02:06 +0000
Subject: resolution of Technical Pan

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Toby
The procedure you describe is not in fact Periodic Acid-Schiff but
the Feulgen reaction, for visualization of DNA in nuclei and
mitochondria.

The PAS reaction uses periodic acid or sodium meta-periodate
(typically 1% solution, ~10min) to oxidise the vicinal diols of some
polysaccharides which are then stained with pararosaniline Schiff's
reagent. It is not usually necessary to use metabisulphite in the
wash water. The detailed procedure is also described in O'Brien &
McCully 1981.

PAS (and Feulgen) may be viewed in a brightfield microscope or by
fluorescence microscopy with a "rhodamine" filter set. An
alternative fluorescent dye which works well in a
fluorescence pseudo Schiff procedure is Lucifer Yellow CH, which
needs blue excitation (FITC filter set).

Chris


Date sent: Mon, 5 Jun 2000 12:29:02 +0930 (CST)
} From: Toby Knight {tknight-at-waite.adelaide.edu.au}
Send reply to: Toby Knight {tknight-at-waite.adelaide.edu.au}


I just came across the following pearl in a 1982 Kodak publication,
number PDS 61 "Selecting film from Kodak for photomicrography".

"The (Technical Pan) 2415 film has very low granularity and is
capable of very great enlargement. In some cases the resolving
power may be beyond that of the microscope image"

How do they know!
Answers on a postcard please ....

Chris
=====================================================================
DR CHRIS JEFFREE
BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
UNIVERSITY OF EDINBURGH
Daniel Rutherford Building
King's Buildings, Mayfield Road
EDINBURGH, EH9 3JH, Scotland, UK
Tel. #44 131 650 5345
FAX. #44 131 650 6563
Mobile 0410 585 401
email c.jeffree-at-ed.ac.uk
SEM / TEM bookings sem-at-ed.ac.uk
=====================================================================




From: Charlesworth, Jon :      charlesworth.jon-at-mayo.edu
Date: Mon, 5 Jun 2000 10:06:40 -0500
Subject: Technician Position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Electron Microscopy Core Facility at Mayo Clinic

The Electron Microscopy Core Facility at Mayo Clinic in
Rochester, MN has an opening for a Biomedical EM technologist to support
both clinical and research projects. The laboratory offers expertise to
collaborative projects that involve transmission and scanning electron
microscopy. The laboratory is well equipped and has a history of excellent
productivity and adequate funding.

The successful candidate for this position will possess at
least a bachelor's degree in Biology with experience in histology and/or
electron microscopy. Additional courses or experience in Immunology, Cell
Biology, and Digital Imaging is desirable. Operating knowledge of
transmission and scanning electron microscopes is preferred. The applicant
must have excellent communicative skills and the ability to work well with a
variety of personalities.

The EM Technologist interacts with all laboratory users in
order to accomplish specific research and clinical goals with respect to
electron microscopy procedures. Duties include: All aspects of specimen
preparation for a variety of biomedical samples for TEM and SEM, operation
of TEM and SEM, darkroom developing and printing, digital image capture, and
reporting. The technologist will also perform advanced research procedures
including immunoelectron microscopy, x-ray microanalysis, and microwave
processing.

Mayo offers a competitive salary and benefits package. If
interested, please submit a cover letter and resume referencing job posting
#00-0002365 to:

Jill Kelly
Mayo Medical Center
Human Resources-OE 1
Rochester, MN 55905
Fax: 507-284-1445
Email: kelly.jill-at-mayo.edu



Jon Charlesworth
Coordinator
Electron Microscopy Core Facility
1426 Gugg
X4-3148




From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Mon, 5 Jun 2000 09:10:09 -0600
Subject: RE: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jim wrote:

'I have two replies and Mike my well have four replies to these.'

I'll try. But I agree with you that we should let this thread die. jim,
it seems you're a bit angy at me. If I unintentionally stepped on your
toe I apologize. I did not expect that we would agree 100%, as you are
selling film and I am involved in digital systems. I just hope that some
other readers found some of this useful. I don't consider myself a
"digital fanatic". This film vs. digital issue has many more facets,
some of which we did not even touch, and which can be just as
entertaining.

Jim wrote:

'Mike reinforced and strengthened that argument, finishing with the note
that he is glad for film when looking at prints by Ansel Adams.'

Just for the record: I usually do not take a magnifying glass to photos.
I mentioned Adams because I like his pictures. If he had taken them with
a digital camera I would have liked them just as much. I also like some
modern art paintings. That does not mean we should start drawing what we
see in the microscopes rather than taking pictures ;-)

Jim wrote:

'Incidentally, from the outset I cited increased "enlargability" to
obtain high
resolution TEM'

Well, you said 'When great enlargements are required film is superior'.
Perhaps I misunderstood. I thought, the term great enlargements referred
to the microscope and meant 'small details', which you can take easier
with a digital camera (no time delay between seeing something and taking
an image, no mechanical vibration due to film movement, etc.). I have
said many times before that film may be better if high resolution AND
large field of view is required.

Jim wrote:

'in any case: Show Sergey!'

I'm in back-channel correspondence with him.



Michael

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: jim [mailto:jim-at-proscitech.com.au]
Sent: Sunday, June 04, 2000 2:34 AM
To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'


I have two replies and Mike my well have four replies to these.
1. There has been a parallel discussion concerning resolution of
film versus
digital images. Clearly in raw power digital cannot compete since film
has
multi gigabyte capacity. I added to that thread that what matters is:
does
digital have enough power and that frequently it would. Mike reinforced
and
strengthened that argument, finishing with the note that he is glad for
film
when looking at prints by Ansel Adams.
(Ansel Adams until about 30 years ago carted for decades large format
cameras
through US National Parks, especially Yosemite, producing fantastic
landscape
photographs and books)
Adams' limited edition prints were contacts of 5x7 and 8x10" inch sheet
film,
probably rated at 400 ISO. The line resolution of such prints much
exceeds our
eyes' resolution, but still results in superior gradation and detail.
TEM film,
even when much enlarged has such details too.
Why Mike, should we accept 2.5 Mb? That is a splendid file size for SEM,
and
because of the limited enlargability of light microscopy (concrete
ceiling due
to wavelengths) it's reasonable, but minimal for light microscopy. TEM
can do
and deserves better.

Incidentally, from the outset I cited increased "enlargability" to
obtain high
resolution TEM, as one of films major advantages, since greater depths
of field
at moderate powers makes high powers through photo enlarging a desirable

technique. The small additional magnification yielded by placing a
digital
camera lower in the column does not compensate. So greater
"enlargability" of
digitals would be desirable, but is limited by pixel size.

2 The thread was initiated by Sergey. He had problems visualising
certain
specimens in dark field. The use of his "beaut" digital TEM camera made
things
worse. I pointed out that the shorter exposure reduced the number of
electrons
forming the image, hence more noise. I believe that a good part of
Mike's case
will be settled in his favour when we hear "Eureka" from Sergey's lab. I
don't
doubt that much more can be done with digital now and that further
improvements
are on the way. Mike's "solution" may well be possible, but I don't
believe its
a snap; in any case: Show Sergey!
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Saturday, June 03, 2000 2:58 AM, Mike Bode [SMTP:mb-at-Soft-Imaging.com]
wrote:
}
} Well, let's see:
}
} Jim wrote:
}
} Mike Bode would use multiple digital
} exposures. The exposures could be layered and combined into one
superior
} image.
} This image would be made up of more pixel and is formed by more
} electrons and
} so would be noise-free and hence could be further enlarged then
} otherwise
} possible. Perhaps.
}
} No, I did not talk about further enlargements. All I wanted to say is,
} that a more noise-free image can be achieved by adding multiple
images,
} and that this also to some extent helps with drift of the sample
during
} acquisition.
}
} Jim wrote:
}
} How much time is required between exposures to transfer a minimum 10mb
} image
} per exposure?
}
} How did you arrive at 10 MB? A 1280x1024 image with 16 bit pixel
} information is about 2.5 MB (uncompressed). We acquire about 10 of
those
} per second and transfer them across the PC bus to the display. Putting
} them on them into Memory might add a few tenth of a second. Writing to
} HD can be done after all images are acquired.
}
} Jim wrote:
}
} What would be the total time from focusing to the last exposure?
} What about Z-drift in the interim requiring objective changes
}
} Why would we have to worry about that, if we don't have to worry about
} that when taking the image on film? In fact, we could take care of
this
} by looking at the image between exposures and correct for z-drift.
} However, as you said, that would add to the overall time and exposure.
I
} was comparing a normal dark field image taken on film at 8 seconds
with
} acquiring the same image on a "too sensitive" CCD camera by adding up
10
} consecutive .8 second images. Why would the sample drift (in x, y or
z)
} substantially more in 8+delta seconds than in 8?
}
} Jim wrote:
}
} what about the total cost of this additional get-up
}
} That of course depends on the microscope and there is no general
answer.
} For example on a LEO 912 I believe the blanker is standard. The
} additional cost to use an acquisition scheme like this with our
software
} is $0 plus perhaps a bit of time to write a small macro. On other
} microscopes one might have to add a beam blanker and perhaps a control
} mechanism for the beam blanker. But I would guess, that this cost is
not
} very high. All modern microscopes are computer controller anyway, so
it
} is most likely just a control command that needs to be sent to the
} microscope over a serial port if the beam blanker is installed. Piece
of
} cake.
}
} Jim wrote:
}
} The mind boggles at a through focus series.
}
} You're right here. But I don't think we were talking about
through-focus
} series. Incidentally, we do through-focus series on light microscopes
} and reconstruction routinely. Takes a few images at different focus
(or
} for a light microscope: stage) settings. The rest is done off-line.
} Takes maybe a couple of minutes for about 20 images of about 1kx1k. I
} agree that TEM is different here and much more complicated due to the
} complicated Contrast Transfer Function. However, this could in
principle
} be sorted out.
}
} Jim wrote:
}
} Again, I don't doubt that there is now a large place for digital in
TEM,
} but
} its no panacea.
}
} I also agree with you on that one. But using the additional computer
} possibilities of digital imaging might take you further than expected.
}
} Michael
}
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 1675 Carr St., #105N
} Lakewood, CO 80215
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} } From: jim [mailto:jim-at-proscitech.com.au]
} Sent: Friday, June 02, 2000 5:15 AM
} To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} Cc: 'jim-at-proscitech.com.au'
} Subject: RE: Film vs Digital
}
}
} The world is full of possible solutions, but are they practical.?
} To produce high-resolution, dark-field or any others TEM images that
} require
} more electrons to form a clear image, Mike Bode would use multiple
} digital
} exposures. The exposures could be layered and combined into one
superior
} image.
} This image would be made up of more pixel and is formed by more
} electrons and
} so would be noise-free and hence could be further enlarged then
} otherwise
} possible. Perhaps.
} Beam blanking would largely save the specimen from beam damage and
drift
} could
} be compensated for by matching up the digitals. Great.
} How much time is required between exposures to transfer a minimum 10mb
} image
} per exposure? What would be the total time from focusing to the last
} exposure?
} What about Z-drift in the interim requiring objective changes and what
} about
} the total cost of this additional get-up. The mind boggles at a
through
} focus
} series.
} When pushing the limits a piece of film seems more effective, cheaper
} and fa
} ster.
} Again, I don't doubt that there is now a large place for digital in
TEM,
} but
} its no panacea.
} Cheers
} Jim Darley
} ProSciTech Microscopy PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} www.proscitech.com
}
} On Friday, June 02, 2000 1:24 AM, Mike Bode [SMTP:mb-at-soft-imaging.com]
} wrote:
} }
} } No, it's not really a problem. It's been done with low density
} } microscopy all the time. Granted, there are some technical aspects
to
} be
} } overcome, but (and I can only speak for ourselves) we have done that
} on
} } a number of microscopes. You are of course correct, that 10 images
at
} } 0.8 seconds take longer than 8 seconds as the image has to be
} } transferred, etc. BUT: that's what beam blankers are for. It is
pretty
} } straightforward to take an image at 0.8 seconds, then blank the beam
} } very quickly before taking the next image. That way you get pretty
} close
} } to the 8 sec total exposure. If there is no beam blanker on the
} } microscope, in most cases it can be added.
} }
} } I am not sure what you mean by "too sensitive". The cameras are
} usually
} } constructed so that 1 electron from the beam creates between a few
} tenth
} } to a few counts (these are all statistical data, of course). The
well
} } width divided by this sensitivity then determines, how many primary
} } electrons are needed to fully expose one pixel. For example, if the
} well
} } width is 50,000 electrons and the sensitivity is 1 count/electron,
one
} } needs 50,000 primary electrons to fill the well. This translates
into
} } roughly a 0.4% statistical error.
} }
} } } From a practical standpoint: You can take images with most cameras
} when
} } the exposure meter on the microscope reads a couple of seconds
without
} } overexposing the camera. On the other hand, you can reduce the
} intensity
} } of the beam until you see single electron events.
} }
} } The one area where CCD cameras may be too sensitive is diffraction.
} The
} } normally huge intensity in the transmitted beam often leads to
} } saturation. In CCDs this can lead to blooming (the intensity spills
} over
} } into neighboring pixels). This can be taken care of with special
chips
} } that have anti-blooming features, but this usually has some other
} } drawbacks. Again, this can also be overcome somewhat with multiple
} } exposures. Film behaves more civilized here, as it simply stops
} } responding to the electrons, but this makes film more or less
useless
} } for quantitative measurements of diffraction patterns. I have done
} } diffraction with CCDs many times and though it does require some
} } tweaking, one can get very good results from them.
} }
} } Michael Bode, Ph.D.
} } Soft Imaging System Corp.
} } 1675 Carr St., #105N
} } Lakewood, CO 80215
} } ===================================
} } phone: (888) FIND SIS
} } (303) 234-9270
} } fax: (303) 234-9271
} } email: mailto:info-at-soft-imaging.com
} } web: http://www.soft-imaging.com
} } ===================================
} }
} }
} }
} } -----Original Message-----
} } From: jim [mailto:jim-at-proscitech.com.au]
} } Sent: Wednesday, May 31, 2000 9:37 PM
} } To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} } Subject: RE: Film vs Digital
} }
} }
} }
}
------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
}
-----------------------------------------------------------------------.
} }
} }
} } Yes, it could be done "in theory". Somebody would need to figure out
} the
} }
} } software and perhaps modify the hardware. Then we would find that
the
} } total
} } exposure of the specimen to the electron beam maybe a muliple of the
} } film's
} } exposure. Afterall, an 8 sec film exposure would not amount in
digital
} } to
} } 10x0.8, but we would require considerable time in between exposures.
} } Since the
} } problems in the discussed circumstances are specimen movement and
beam
} } damage,
} } it seems that taking multiple exposures is a poor option.
} }
} } Digital cameras are for some situation too sensitive to electron
} } exposure.
} } Cutting back on electrons is no option since its the electrons that
} form
} } the
} } image in the first instance.
} } Much easier in light microscopy . . . insert a neutral density
} filter.
} } Cheers
} } Jim Darley
} } ProSciTech Microscopy PLUS
} } PO Box 111, Thuringowa QLD 4817 Australia
} } Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} } Great microscopy catalogue, 500 Links, MSDS, User Notes
} } www.proscitech.com
} }





From: Nestor J. Zaluzec :      zaluzec-at-sparc5.microscopy.com
Date: Mon, 5 Jun 2000 10:55:39 -0500
Subject: Re: Administrivia: Resolving power: - Clarification from Nestor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jonathon
Low kV is preferred for some biological applications chiefly
because biologists experience difficulty getting enough contrast out
of their specimens. As Dr Alves de Matos rightly says, low kV is
more damaging to specimens than high kV, because the
penetrating power of low-kV electrons is lower and therefore more
beam energy is lost to the specimen. (We think of this exactly the
other way round for SEM, but that's a story for another day!).

The main reason for the apparent shift in biologist's attitudes is that
many are now less concerned with cell and tissue-level
ultrastructure and increasingly interested in imaging
macromolecules. 200kV or 300kV machines are preferred for cryo-
microscopy of macromolecules in vitrified ice because the
resolution is a lot better than at 80 or 120kV, and also because
200kV causes less specimen damage.

However, the cost of resolution either from higher kVs or the
shorter working distance objective lenses favoured by materials
scientists is poorer image contrast. Philips therefore supply
"Biotwin" versions of some of their TEM range which use a long
focal length objective which trading a little resolution for a little
more contrast. This optical configuration is coincidentally very good
for thickish specimens such as ultrathin biological sections where
the ultimate resolution of the machine is pretty much academic
anyway.

Images of thick (#100nm) specimens such as FIB sections of
microelectronic devices seem crisper and more contrasty in our
CM120 Biotwin than in a 200KV machine.
Chris

} From: "A.P.Alves de Matos" {apmatos-at-ip.pt}
To: {Microscopy-at-sparc5.microscopy.com}


Colleagues....

Some of you have misinterpreted my earlier message.
Please feel free to carry on the discussion on film/digitization etc..

I only asked that the "tangental thread" centered about
complaints on a specific product be taken off-line before
any fire and brimstone starts up between a manufacturer
and clients. This is not the forum for that type of feedback.

The previous dicussion should certainly continue as long
as necessary.

Nestor
Your Friendly Neighborhood SysOp



}
} Nestor,
}
} It would be very unfortunate to squelch this important and central debate
} concerning not only the corresponding resolution comparisons but also
} differances in performance due to noise and greyscale depth. New methods of
} enhansing resolution need to be critiqued as well.
}
} Dave Barnard
}
} Wadsworth Center HVEM
} NYS Dept. Health
} Albany NY







From: Sha Zhu :      szhu-at-engin.umich.edu
Date: Mon, 5 Jun 2000 16:00:52 -0400 (EDT)
Subject: TEM: looking for T-tool

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




We are looking for the "T-tool" - a kind of fixture for TEM sample
polishing. It is from T&T group in USA. We need to know the email
address or telephone number of T&T group.

Any information about the T&T group or the T-tool will be highly
appreciated!

Thanks a lot!

szhu






From: David Joswiak :      joswiak-at-orca.astro.washington.edu
Date: Mon, 5 Jun 2000 13:30:28 -0700 (PDT)
Subject: Wanted: Used JEOL 2010 TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are looking to purchase a used JEOL 2010 TEM (LaB6, not FE) in good
working condition. I would appreciate replies from anyone who knows if a
JEOL 2010 is available or will become available in the next 6 months or
so. A STEM unit would increase our interest, but isn't strictly
necessary.

Also, we will be looking for a buyer for our JEOL 1200EX STEM which has
been under service contract from day 1. It has been a highly reliable
microscope and has given years of satisfactory performance.

Dave Joswiak
Research Scientist
Dept. of Astronomy, 351580
University of Washington
Seattle, WA 98195
(206)543-7702
joswiak-at-astro.washington.edu







From: John J. Bozzola :      bozzola-at-siu.edu
Date: Mon, 5 Jun 2000 16:14:59 -0500
Subject: SEM: softening of beetle elytra

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am forwarding this request on behalf of Mr. Edoardo Nolfo. If
anyone could be of assistance please send the information to him
directly or (if you prefer) I will forward it to him.

Thank you.

J.B.


} From: Edoardo Nolfo {edoardo.nolfo-at-christ-church.oxford.ac.uk}

} I am a student at the University of Oxford and am writing to ask for some
} advice. I am currently working on multilayer reflectors in beetles; I plan to
} use S.E.M. and T.E.M. to elucidate the fine structure of these reflectors.
} Apparently the elytra of the beetles must be sliced very thinly for S.E.M.
} analysis. My problem lies in the fact that the elytra are extremely
} hard, which
} means they cannot be sliced very thinly. Do you have any advice to
} offer on how
} to soften the elytra for this purpose? Perhaps there is a chemical
} that is used
} to make samples softer, or a standard technique used to soften very hard
} samples. I would be extremely grateful to you if you could help me! Please do
} not hesitate to contact me if you require any more information.
}
} Let me thank you in advance for your help; I look forward to hearing from you
} soon.

####################################################################
John J. Bozzola, Ph.D., Director
Micro-Imaging and Analysis Center
750 Communications Drive - MC 4402
Southern Illinois University
Carbondale, IL 62901 U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################




From: John Nailon :      mmjnailo-at-dingo.cc.uq.edu.au
Date: Sun, 04 Jun 2000 08:54:05 +1000
Subject: Re: resolution of Technical Pan

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chris,
This is because Kodak use EMs to measure the grain size of their silver
halide grains. Tech Pan has a very small grain size (slow film, low
ISO/ASA number) and is therefore capable of greater enlargement than
film with a larger grains. The larger the grain the faster the film
(higher ISO/ASA number).
Regards
JVN
JVN

Chris Jeffree wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} I just came across the following pearl in a 1982 Kodak publication,
} number PDS 61 "Selecting film from Kodak for photomicrography".
}
} "The (Technical Pan) 2415 film has very low granularity and is
} capable of very great enlargement. In some cases the resolving
} power may be beyond that of the microscope image"
}
} How do they know!
} Answers on a postcard please ....
}
} Chris
} =====================================================================
} DR CHRIS JEFFREE
} BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
} UNIVERSITY OF EDINBURGH
} Daniel Rutherford Building
} King's Buildings, Mayfield Road
} EDINBURGH, EH9 3JH, Scotland, UK
} Tel. #44 131 650 5345
} FAX. #44 131 650 6563
} Mobile 0410 585 401
} email c.jeffree-at-ed.ac.uk
} SEM / TEM bookings sem-at-ed.ac.uk
} =====================================================================

--
****************************************************
John V Nailon
Operations Manager
Centre for Microscopy and Microanalysis
The University of Queensland
St. Lucia Queensland 4072
Phone: +61-7-3365-4214
Fax: +61-7-3365-4422
WWW: http://www.uq.edu.au/nanoworld/allstaff.html#Nailon
****************************************************




From: David Henriks :      Henriks-at-CompuServe.COM
Date: Mon, 5 Jun 2000 21:32:46 -0400
Subject: TEM: looking for T-tool

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dr. Zhu:

The "T" tool is a variation of the well known Tripod Polisher¨. We are the
manufacturers of the Tripod Polisher¨ and also manufacture the more compact
BiPod Polisher. The BiPod polisher offers the smaller size as found on the
T-tool while still incorporating our many years of experience in SEM and
TEM cross section polishing as well as our expertise in precision
machining. I know this doesn't help much in locating the T-tool, but it
does offer you the information you need to acquire a tool that will
effectively meet your requirements. If you would like additional
information on these tools, please feel free to contact me.

Best regards-

David
Writing at 6:24:03 PM on 06/05/2000

***************************************************************************
************************

David Henriks
Vice President TEL: 800-728-2233
(toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499
San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com

***************************************************************************
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.



Message text written by Sha Zhu
} -----------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America




We are looking for the "T-tool" - a kind of fixture for TEM sample
polishing. It is from T&T group in USA. We need to know the email
address or telephone number of T&T group.

Any information about the T&T group or the T-tool will be highly
appreciated!

Thanks a lot!


szhu

{




From: Khanh Tran :      khanh-at-deakin.edu.au
Date: Tue, 06 Jun 2000 14:01:40 +1000
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello ,

please subscribe
Khanh Tran
Deakin University
662 Blackburn Road
CLAYTON, VIC. 3168
AUSTRALIA




From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Mon, 5 Jun 2000 09:10:09 -0600
Subject: RE: Film vs Digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Jim wrote:

'I have two replies and Mike my well have four replies to these.'

I'll try. But I agree with you that we should let this thread die. jim,
it seems you're a bit angy at me. If I unintentionally stepped on your
toe I apologize. I did not expect that we would agree 100%, as you are
selling film and I am involved in digital systems. I just hope that some
other readers found some of this useful. I don't consider myself a
"digital fanatic". This film vs. digital issue has many more facets,
some of which we did not even touch, and which can be just as
entertaining.

Jim wrote:

'Mike reinforced and strengthened that argument, finishing with the note
that he is glad for film when looking at prints by Ansel Adams.'

Just for the record: I usually do not take a magnifying glass to photos.
I mentioned Adams because I like his pictures. If he had taken them with
a digital camera I would have liked them just as much. I also like some
modern art paintings. That does not mean we should start drawing what we
see in the microscopes rather than taking pictures ;-)

Jim wrote:

'Incidentally, from the outset I cited increased "enlargability" to
obtain high
resolution TEM'

Well, you said 'When great enlargements are required film is superior'.
Perhaps I misunderstood. I thought, the term great enlargements referred
to the microscope and meant 'small details', which you can take easier
with a digital camera (no time delay between seeing something and taking
an image, no mechanical vibration due to film movement, etc.). I have
said many times before that film may be better if high resolution AND
large field of view is required.

Jim wrote:

'in any case: Show Sergey!'

I'm in back-channel correspondence with him.



Michael

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: jim [mailto:jim-at-proscitech.com.au]
Sent: Sunday, June 04, 2000 2:34 AM
To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'


I have two replies and Mike my well have four replies to these.
1. There has been a parallel discussion concerning resolution of
film versus
digital images. Clearly in raw power digital cannot compete since film
has
multi gigabyte capacity. I added to that thread that what matters is:
does
digital have enough power and that frequently it would. Mike reinforced
and
strengthened that argument, finishing with the note that he is glad for
film
when looking at prints by Ansel Adams.
(Ansel Adams until about 30 years ago carted for decades large format
cameras
through US National Parks, especially Yosemite, producing fantastic
landscape
photographs and books)
Adams' limited edition prints were contacts of 5x7 and 8x10" inch sheet
film,
probably rated at 400 ISO. The line resolution of such prints much
exceeds our
eyes' resolution, but still results in superior gradation and detail.
TEM film,
even when much enlarged has such details too.
Why Mike, should we accept 2.5 Mb? That is a splendid file size for SEM,
and
because of the limited enlargability of light microscopy (concrete
ceiling due
to wavelengths) it's reasonable, but minimal for light microscopy. TEM
can do
and deserves better.

Incidentally, from the outset I cited increased "enlargability" to
obtain high
resolution TEM, as one of films major advantages, since greater depths
of field
at moderate powers makes high powers through photo enlarging a desirable

technique. The small additional magnification yielded by placing a
digital
camera lower in the column does not compensate. So greater
"enlargability" of
digitals would be desirable, but is limited by pixel size.

2 The thread was initiated by Sergey. He had problems visualising
certain
specimens in dark field. The use of his "beaut" digital TEM camera made
things
worse. I pointed out that the shorter exposure reduced the number of
electrons
forming the image, hence more noise. I believe that a good part of
Mike's case
will be settled in his favour when we hear "Eureka" from Sergey's lab. I
don't
doubt that much more can be done with digital now and that further
improvements
are on the way. Mike's "solution" may well be possible, but I don't
believe its
a snap; in any case: Show Sergey!
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Saturday, June 03, 2000 2:58 AM, Mike Bode [SMTP:mb-at-Soft-Imaging.com]
wrote:
}
} Well, let's see:
}
} Jim wrote:
}
} Mike Bode would use multiple digital
} exposures. The exposures could be layered and combined into one
superior
} image.
} This image would be made up of more pixel and is formed by more
} electrons and
} so would be noise-free and hence could be further enlarged then
} otherwise
} possible. Perhaps.
}
} No, I did not talk about further enlargements. All I wanted to say is,
} that a more noise-free image can be achieved by adding multiple
images,
} and that this also to some extent helps with drift of the sample
during
} acquisition.
}
} Jim wrote:
}
} How much time is required between exposures to transfer a minimum 10mb
} image
} per exposure?
}
} How did you arrive at 10 MB? A 1280x1024 image with 16 bit pixel
} information is about 2.5 MB (uncompressed). We acquire about 10 of
those
} per second and transfer them across the PC bus to the display. Putting
} them on them into Memory might add a few tenth of a second. Writing to
} HD can be done after all images are acquired.
}
} Jim wrote:
}
} What would be the total time from focusing to the last exposure?
} What about Z-drift in the interim requiring objective changes
}
} Why would we have to worry about that, if we don't have to worry about
} that when taking the image on film? In fact, we could take care of
this
} by looking at the image between exposures and correct for z-drift.
} However, as you said, that would add to the overall time and exposure.
I
} was comparing a normal dark field image taken on film at 8 seconds
with
} acquiring the same image on a "too sensitive" CCD camera by adding up
10
} consecutive .8 second images. Why would the sample drift (in x, y or
z)
} substantially more in 8+delta seconds than in 8?
}
} Jim wrote:
}
} what about the total cost of this additional get-up
}
} That of course depends on the microscope and there is no general
answer.
} For example on a LEO 912 I believe the blanker is standard. The
} additional cost to use an acquisition scheme like this with our
software
} is $0 plus perhaps a bit of time to write a small macro. On other
} microscopes one might have to add a beam blanker and perhaps a control
} mechanism for the beam blanker. But I would guess, that this cost is
not
} very high. All modern microscopes are computer controller anyway, so
it
} is most likely just a control command that needs to be sent to the
} microscope over a serial port if the beam blanker is installed. Piece
of
} cake.
}
} Jim wrote:
}
} The mind boggles at a through focus series.
}
} You're right here. But I don't think we were talking about
through-focus
} series. Incidentally, we do through-focus series on light microscopes
} and reconstruction routinely. Takes a few images at different focus
(or
} for a light microscope: stage) settings. The rest is done off-line.
} Takes maybe a couple of minutes for about 20 images of about 1kx1k. I
} agree that TEM is different here and much more complicated due to the
} complicated Contrast Transfer Function. However, this could in
principle
} be sorted out.
}
} Jim wrote:
}
} Again, I don't doubt that there is now a large place for digital in
TEM,
} but
} its no panacea.
}
} I also agree with you on that one. But using the additional computer
} possibilities of digital imaging might take you further than expected.
}
} Michael
}
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 1675 Carr St., #105N
} Lakewood, CO 80215
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} } From: jim [mailto:jim-at-proscitech.com.au]
} Sent: Friday, June 02, 2000 5:15 AM
} To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} Cc: 'jim-at-proscitech.com.au'
} Subject: RE: Film vs Digital
}
}
} The world is full of possible solutions, but are they practical.?
} To produce high-resolution, dark-field or any others TEM images that
} require
} more electrons to form a clear image, Mike Bode would use multiple
} digital
} exposures. The exposures could be layered and combined into one
superior
} image.
} This image would be made up of more pixel and is formed by more
} electrons and
} so would be noise-free and hence could be further enlarged then
} otherwise
} possible. Perhaps.
} Beam blanking would largely save the specimen from beam damage and
drift
} could
} be compensated for by matching up the digitals. Great.
} How much time is required between exposures to transfer a minimum 10mb
} image
} per exposure? What would be the total time from focusing to the last
} exposure?
} What about Z-drift in the interim requiring objective changes and what
} about
} the total cost of this additional get-up. The mind boggles at a
through
} focus
} series.
} When pushing the limits a piece of film seems more effective, cheaper
} and fa
} ster.
} Again, I don't doubt that there is now a large place for digital in
TEM,
} but
} its no panacea.
} Cheers
} Jim Darley
} ProSciTech Microscopy PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} www.proscitech.com
}
} On Friday, June 02, 2000 1:24 AM, Mike Bode [SMTP:mb-at-soft-imaging.com]
} wrote:
} }
} } No, it's not really a problem. It's been done with low density
} } microscopy all the time. Granted, there are some technical aspects
to
} be
} } overcome, but (and I can only speak for ourselves) we have done that
} on
} } a number of microscopes. You are of course correct, that 10 images
at
} } 0.8 seconds take longer than 8 seconds as the image has to be
} } transferred, etc. BUT: that's what beam blankers are for. It is
pretty
} } straightforward to take an image at 0.8 seconds, then blank the beam
} } very quickly before taking the next image. That way you get pretty
} close
} } to the 8 sec total exposure. If there is no beam blanker on the
} } microscope, in most cases it can be added.
} }
} } I am not sure what you mean by "too sensitive". The cameras are
} usually
} } constructed so that 1 electron from the beam creates between a few
} tenth
} } to a few counts (these are all statistical data, of course). The
well
} } width divided by this sensitivity then determines, how many primary
} } electrons are needed to fully expose one pixel. For example, if the
} well
} } width is 50,000 electrons and the sensitivity is 1 count/electron,
one
} } needs 50,000 primary electrons to fill the well. This translates
into
} } roughly a 0.4% statistical error.
} }
} } } From a practical standpoint: You can take images with most cameras
} when
} } the exposure meter on the microscope reads a couple of seconds
without
} } overexposing the camera. On the other hand, you can reduce the
} intensity
} } of the beam until you see single electron events.
} }
} } The one area where CCD cameras may be too sensitive is diffraction.
} The
} } normally huge intensity in the transmitted beam often leads to
} } saturation. In CCDs this can lead to blooming (the intensity spills
} over
} } into neighboring pixels). This can be taken care of with special
chips
} } that have anti-blooming features, but this usually has some other
} } drawbacks. Again, this can also be overcome somewhat with multiple
} } exposures. Film behaves more civilized here, as it simply stops
} } responding to the electrons, but this makes film more or less
useless
} } for quantitative measurements of diffraction patterns. I have done
} } diffraction with CCDs many times and though it does require some
} } tweaking, one can get very good results from them.
} }
} } Michael Bode, Ph.D.
} } Soft Imaging System Corp.
} } 1675 Carr St., #105N
} } Lakewood, CO 80215
} } ===================================
} } phone: (888) FIND SIS
} } (303) 234-9270
} } fax: (303) 234-9271
} } email: mailto:info-at-soft-imaging.com
} } web: http://www.soft-imaging.com
} } ===================================
} }
} }
} }
} } -----Original Message-----
} } From: jim [mailto:jim-at-proscitech.com.au]
} } Sent: Wednesday, May 31, 2000 9:37 PM
} } To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} } Subject: RE: Film vs Digital
} }
} }
} }
}
------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
}
-----------------------------------------------------------------------.
} }
} }
} } Yes, it could be done "in theory". Somebody would need to figure out
} the
} }
} } software and perhaps modify the hardware. Then we would find that
the
} } total
} } exposure of the specimen to the electron beam maybe a muliple of the
} } film's
} } exposure. Afterall, an 8 sec film exposure would not amount in
digital
} } to
} } 10x0.8, but we would require considerable time in between exposures.
} } Since the
} } problems in the discussed circumstances are specimen movement and
beam
} } damage,
} } it seems that taking multiple exposures is a poor option.
} }
} } Digital cameras are for some situation too sensitive to electron
} } exposure.
} } Cutting back on electrons is no option since its the electrons that
} form
} } the
} } image in the first instance.
} } Much easier in light microscopy . . . insert a neutral density
} filter.
} } Cheers
} } Jim Darley
} } ProSciTech Microscopy PLUS
} } PO Box 111, Thuringowa QLD 4817 Australia
} } Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} } Great microscopy catalogue, 500 Links, MSDS, User Notes
} } www.proscitech.com
} }






From: Didier Le Thiec :      le_thiec-at-nancy.inra.fr
Date: Tue, 6 Jun 2000 09:09:33 -0500
Subject: leaves

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Jonathon
Low kV is preferred for some biological applications chiefly
because biologists experience difficulty getting enough contrast out
of their specimens. As Dr Alves de Matos rightly says, low kV is
more damaging to specimens than high kV, because the
penetrating power of low-kV electrons is lower and therefore more
beam energy is lost to the specimen. (We think of this exactly the
other way round for SEM, but that's a story for another day!).

The main reason for the apparent shift in biologist's attitudes is that
many are now less concerned with cell and tissue-level
ultrastructure and increasingly interested in imaging
macromolecules. 200kV or 300kV machines are preferred for cryo-
microscopy of macromolecules in vitrified ice because the
resolution is a lot better than at 80 or 120kV, and also because
200kV causes less specimen damage.

However, the cost of resolution either from higher kVs or the
shorter working distance objective lenses favoured by materials
scientists is poorer image contrast. Philips therefore supply
"Biotwin" versions of some of their TEM range which use a long
focal length objective which trading a little resolution for a little
more contrast. This optical configuration is coincidentally very good
for thickish specimens such as ultrathin biological sections where
the ultimate resolution of the machine is pretty much academic
anyway.

Images of thick (#100nm) specimens such as FIB sections of
microelectronic devices seem crisper and more contrasty in our
CM120 Biotwin than in a 200KV machine.
Chris

} From: "A.P.Alves de Matos" {apmatos-at-ip.pt}
To: {Microscopy-at-sparc5.microscopy.com}


Dear all,

1) I am working on oak trees, and I would like to describe the
ultrastucture of leaves; also I am looking for a protocol (very detailed)
to do it.
2) On oak leaves I would like to localize the superoxyde
dismutase(s) enzymes by immunocytochemistry, also I am looking for a
protocol (very detailed) to do it.

Thanks in advances.
Dr. Didier Le Thiec

--------------------------------------------
Didier Le Thiec
I.N.R.A. Centre de Recherches Forestieres
Unite d'Ecophysiologie Forestiere
Laboratoire de Pollution Atmospherique
54280 Champenoux - France

Tel : 33 (0)3 83 39 40 98
Fax : 33 (0)3 83 39 40 69
E-mail : le_thiec-at-nancy.inra.fr
http://vectra.nancy.inra.fr/pollu/index.htm
-------------------------------------------






From: Rafal Dunin-Borkowski :      rafal.db-at-materials.oxford.ac.uk
Date: Tue, 6 Jun 2000 09:26:36 +0100 (BST)
Subject: FEG/TEM one-day meeting, Monday 3 July 2000

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Monday 3 July 2000

New Developments in FEG/TEM

A joint one-day meeting organised by RMS and EMAG

Department of Materials, Oxford University

Following on from the very successful meeting held in 1999, this will
review the current state-of-the-art of the UK's various FEG/TEM
installations in operation.

The meeting will begin with a special keynote lecture on

Advanced TEM Instrumentation for Solving Critical Problems in Materials
Science given by Professor Manfred Rźhle Max-Planck Institute fźr
Metallforschung, Stuttgart

This lecture will include some news on the new German FEG/TEM project.

There are now ten FEG/TEM facilities in the UK (some with more than one
instrument) and the main programme will consist of a series of invited
presentations from each of the sites, reviewing the latest developments in
instrumentation and covering specialised techniques such as holography,
energy-filtered imaging, spectroscopic imaging, nano-scale analysis, etc.

With JIF, JREI and other special equipment funding now becoming available
it is likely that there will be other sophisticated FEG/TEM installations
in place in the near future. With this in view the meeting will also
include a forward look: what will the next generation of instruments be
like?

Speakers:

John Hutchison (University of Oxford)

Key-note Lecture - Professor Manfred RÄhle (Stuttgart)
Per Bullough (University of Sheffield)
Marin van Heel (Imperial College)
John Berriman (MRC, Cambridge)
Tony Cullis (University of Sheffield)
Ian Jones (University of Birmingham)
Rik Brydson (University of Leeds)
Jeremy Sloan (University of Oxford)
David Cherns (University of Bristol)
Stephen McVitie (Glasgow University)
Stephen Lloyd (University of Cambridge)
John Titchmarsh (University of Oxford)
Paul Midgley (University of Cambridge)


The meeting will conclude with the RMS AGM and Presidential Address.

For further details contact:
Dr Paul Midgley (EMAG): pam33-at-cam.ac.uk or
Dr John Hutchison (RMS): john.hutchison-at-materials.ox.ac.uk





From: Tobias Baskin :      BaskinT-at-missouri.edu
Date: Tue, 6 Jun 2000 11:06:50 -0400
Subject: Re: resolution of Technical Pan

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
Just for the record, Tech Pan is interesting in that you can
vary the grain size over an amzingly wide range by choice of
developer and exposure conditions. You can technidol LC (I think) and
get the teeeny weeeeny grains previously mentioned, or use D19 (I
think) and get really high contrast big grains.

Just my few exposed halides,
Tobias



} Chris,
} This is because Kodak use EMs to measure the grain size of their silver
} halide grains. Tech Pan has a very small grain size (slow film, low
} ISO/ASA number) and is therefore capable of greater enlargement than
} film with a larger grains. The larger the grain the faster the film
} (higher ISO/ASA number).
} Regards
} JVN
} JVN
}
} Chris Jeffree wrote:
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} } I just came across the following pearl in a 1982 Kodak publication,
} } number PDS 61 "Selecting film from Kodak for photomicrography".
} }
} } "The (Technical Pan) 2415 film has very low granularity and is
} } capable of very great enlargement. In some cases the resolving
} } power may be beyond that of the microscope image"
} }
} } How do they know!
} } Answers on a postcard please ....
} }
} } Chris
} } =====================================================================
} } DR CHRIS JEFFREE
} } BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
} } UNIVERSITY OF EDINBURGH
} } Daniel Rutherford Building
} } King's Buildings, Mayfield Road
} } EDINBURGH, EH9 3JH, Scotland, UK
} } Tel. #44 131 650 5345
} } FAX. #44 131 650 6563
} } Mobile 0410 585 401
} } email c.jeffree-at-ed.ac.uk
} } SEM / TEM bookings sem-at-ed.ac.uk
} } =====================================================================
}
} --
} ****************************************************
} John V Nailon
} Operations Manager
} Centre for Microscopy and Microanalysis
} The University of Queensland
} St. Lucia Queensland 4072
} Phone: +61-7-3365-4214
} Fax: +61-7-3365-4422
} WWW: http://www.uq.edu.au/nanoworld/allstaff.html#Nailon
} ****************************************************

_ ____ ^ __ ____ Tobias I. Baskin
/ \ / / \ / \ \ University of Missouri
/ | / / \ \ \ Biological Sciences
/___/ /__ /___ \ \ \__ 109 Tucker Hall
/ / / \ \ \ Columbia, MO 65211-7400 USA
/ / / \ \ \ voice: 573-882-0173
/ /____ / \ \__/ \____ fax: 573-882-0123




From: Ronald Vane :      RVaneXEI-at-concentric.net
Date: Wednesday, May 24, 2000 10:58 AM
Subject: Re: species analysis in real time

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


MSA Listers:

Two weeks ago I posted a short message about the Ocean Optics spectrometer I
had just received for doing Plasma diagnostics. I tried it and have found it
very useful for examining a plasma cleaning process and optimizing operating
conditions. I purchased a USB 2000 model that plugs into a USB port of a PC
for less than $3000. Ocean Optics, 727-733-2447, www.OceanOptics.com

Ronald Vane
XEI Scientific
NEW WEB SITE! www.SEMCLEAN.com


-----Original Message-----
} From: Ronald Vane {RVaneXEI-at-concentric.net}
To: MSA listserver {Microscopy-at-sparc5.microscopy.com} ; Dr. Gary Gaugler
{gary-at-gaugler.com}





From: =shAf= :      mshaf-at-darkwing.uoregon.edu
Date: Tue, 6 Jun 2000 10:17:46 -0700
Subject: SEM: agency wants SEM images?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My SEM facility has just been approached, and in the interest in
providing what they want for a price based on an average, can I query
the SEM community as to what they charge for commercial release of
their SEM imagery. I'm looking for $/image, but if you charge on any
type of sliding scale, I'd be interested in that too. I believe as
well I shouldn't be undercutting commercial facilities, so I'm
especially interested in these numbers.
Reply to me direct, and I'll respond back to the list with the
average, max and min, and no names.

cheerios, =shAf= :o)

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - mshaf-at-darkwing.uoregon.edu
Geological Science's Electron Probe Facility - University of Oregon
http://epmalab.uoregon.edu/






From: anderron-at-us.ibm.com
Date: Tue, 6 Jun 2000 13:32:30 -0400
Subject: Fluorescence in-situ Hybridization (FISH) Course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This individual contacted the MSA Business Office with the following
request. Please help her if you can, offline. Don't reply back to me.

Thanks,

..snip

Content-Type: text/plain
Content-Disposition: inline

I work for North Carolina State University and we are looking in to
using Fluorescence in-situ Hybridization (FISH) for identification of
bacteria in biofilm. Do you know of any short courses in FISH here in
the USA? I found some out of the country but would prefer to take a
course in USA.

Thanks for your time

Tracey L. Daly {tldaly-at-unity.ncsu.edu}




Ron Anderson, IBM, Hopewell Jct., New York, USA. anderron-at-us.ibm.com

IBM Analytical Services; http://www.chips.ibm.com/services/asg





From: jmkrupp-at-cats.ucsc.edu (Jon Krupp)
Date: Tue, 6 Jun 2000 11:13:51 -0700
Subject: PGT parts for anyone

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi:

I am about to tear the guts out of an old PGT System4 console. I need some
of the electronics, but will toss the rest unless someone would like it.

Jonathan Krupp
Microscopy & Imaging Lab
University of California
Santa Cruz, CA 95064
(831) 459-2477
jmkrupp-at-cats.ucsc.edu






From: donald j marshall :      dmrelion-at-world.std.com
Date: Tue, 6 Jun 2000 15:27:35 -0400 (EDT)
Subject: UV optics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have an associate who is doing cathodoluminescence studies on materials.
He would like to extend his observations into the UV for a limited number of
tests on a a few samples. He is presently limited by the optics of the
microscope and the fiber optics cable connecting the microscope ocular to
the input of the spectrometer. Does anyone have any suggestions for optical
microscopes with UV optics that might be available for short term lease or
loan that he could consider? Objective magnifications of 5X to 10 X and a 5X
or 10X ocular would probably suffice.

Thanks,

Don Marshall


Donald J. Marshall
Relion Industries
P.O. Box 12
Bedford, MA 01730
Ph: 781-275-4695
FAX: 781-271-0252
email dmrelion-at-world.std.com

Cathodoluminescence, mass spectroscopy, electron beam technology


"A weed is a flower out of place."

(Please note: Do not send email with attachments to this address. Instead, send it to donbarlen-at-aol.com. Thank you.)




From: Jane Fagerland :      jane.a.fagerland-at-abbott.com
Date: Tue, 6 Jun 2000 16:22:14 -0500
Subject: Position Opening - Biological Electron Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Department of Microscopy and Microanalysis at Abbott Laboratories is
recruiting an Electron Microscopist for its Biological Microscopy group.
This group provides ultrastructural pathology support for Nonclinical Drug
Safety studies, as well as for other biological microscopy projects, such as
cell screening, virus identification and counting, and immunolabeling.

Requirements for this position include:
a Masters' degree in a biological field, such as cell biology, anatomy, or
zoology
a thorough understanding of mammalian histology and ultrastructure
excellent technical skills including tissue collection at necropsy, tissue
and cell processing for TEM, sectioning, staining, operation of electron
microscopes and related equipment, darkroom procedures
ability to communicate information in technical reports and in oral
presentations

Highly desirable, but not essential, are:
experience in ultrastructural pathology or toxicologic pathology
working knowledge of SEM specimen preparation and instrumentation
working knowledge of immunocytochemistry, in situ hybridization, and other
labeling methods at light or electron microscopic level
familiarity with Good Laboratory Practices

We are looking for a team player with outstanding interpersonal skills and
the ability to adjust readily to rapidly changing priorities and shifting
deadlines. The ability to communicate clearly, both verbally and in writing,
is essential. Careful attention to detail and accuracy are required.

The Department of Microscopy and Microanalysis provides corporate-wide
support in Biological and Materials Microscopy to all divisions of Abbott
Laboratories. The facility houses two TEMs (a Philips CM12 STEM and a LEO
910), two SEMs (a Philips XL30-FEG and AMRAY 1830i), three EDXS systems, a
BioRad confocal scanning laser microscope, several fluorescent and light
microscopes (polarized light, DIC), and a Quantimet Image Analysis system.
We also have an Arcturus PixCell laser capture microdissection system and a
Becton Dickinson FACSCalibur flow cytometer. Microtomes include Reichert
Ultracut E and S ultramicrotomes, RMC 6000 XL cryoultrotome, Microm
histological microtome, and Microm HM500 cryostat.

Please send letters of application and resumes to:

Jane A. Fagerland, Ph.D.
Abbott Laboratories
D45M/AP31
200 Abbott Park RD.
Abbott Park IL 60064-6202

(847) 935-0104 voice
(847) 938-5027 fax
jane.a.fagerland -at-abbott.com




From: Christensen, Kim :      ChristeK-at-whiteoaksemi.com
Date: Tue, 6 Jun 2000 17:42:54 -0400
Subject: Re: T-Tool

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sha Zhu,

The T-Tool is available from:
Precision TEM, Inc.
Santa Clara, CA
Ph: 408-980-8898
E-mail: liya-at-precisiontem.com

----------------------------------------------------
Kim Christensen
Failure Analysis Laboratory
White Oak Semiconductor
6000 Technology Boulevard
Sandston, Virginia 23150
Ph: 804 952 7307
Fax: 804 952 7902
Pager: 1 800 759 8888, pin# 130 3382
----------------------------------------------------




From: guerrier-at-ews.uiuc.edu, GUERRGI-at-mail.northgrum.com ()
Date: Tue, 6 Jun 2000 23:43:28 -0500
Subject: Help? Cleaning Optical Lenses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Email: guerrier-at-ews.uiuc.edu, GUERRGI-at-mail.northgrum.com
Name: Gina Guerrieri

School: University of Illinois


Question: How do you clean a stereo zoom microscope? Specifically, how do
you clean the lenses with out removing them from the system?

---------------------------------------------------------------------------






From: Kristof Kovacs :      kris-at-almos.vein.hu
Date: Wed, 07 Jun 2000 08:37:32 +0200
Subject: Still alive?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,
I'm just curious if something is wrong on my end or I'm unsuscribed against
my will. I did not receive any posting during the past two-three weeks, and
this is impossible in view of previously received 10-20 postings/day.
Is there any explanation?
Kris
Dr. Kristof Kovacs
Associate Professor
President, Hungarian Society for Microscopy
Phone: +36-(88)-421-684
Fax: +36-(88)-328-643
Mailing Address:
University of Veszprem, P.O.Box 158, Veszprem
H-8201 Hungary





From: Ian MacLaren :      maclariz-at-yahoo.co.uk
Date: Wed, 7 Jun 2000 08:27:57 +0100 (BST)
Subject: Ceramic cross-section preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,
I'm currently having a few problems preparing good cross-section specimens
of ceramic thin films on LaAlO3 or NdGaO3 substrates.

I can cut them okay, but thinning them often results in cracking as the
thickness goes below 100 microns. I have a Gatan model 623 disk grinder and
currently have access to SiC abrasive papers to grit sizes of 1500 or 2000,
and also have the Gatan diamond polishing disks and specimen lapping kit.

The problems are as follows:
If we use the 1500 or 2000 grit SiC, we can often get the samples thin
enough, but with poor surface finish which needs to be improved with
dimpling (on both sides), we also risk cracking the sample as SiC papers
often contain imperfections.

Using the Gatan specimen lapping kit, I find that as the disk grinder is
altered to grind off a further 10 microns, the edge of the sample often
catches on the diamond disk and tears some of the diamond coating off,
leaving a lump on the surface which then risks catching the specimen and
cracking it.

I have one possible alternative approach which I used in Sweden last year
(with SiAlON ceramics) involving attaching the sample to a glass slide and
polishing it using diamond spray on non-absorbent paper. I may consider
trying this with these new materials.

Other ideas would be welcome, however, as would suggestions of how to
improve my present method. Please note that I do not have the budget to buy
any expensive new polishing equipment (such as a tripod polisher, for
instance), so the most welcome suggestions would be those that involve
improvements to current techniques, use of different types of polishing
consumables, etc..

I hope to hear several suggestions, both from users and from the companies
who producing polishing and lapping equipment. Why not post them with the
list so we can all benefit from the sharing of experience?

Best wishes

=====
Ian MacLaren
Beijing Laboratory of Electron Microscopy
Chinese Academy of Sciences, P.O. Box 2724
100080 Beijing
China
General Email: ian.maclaren-at-physics.org
Work (esp. large attachments): maclaren-at-image.blem.ac.cn

____________________________________________________________
Do You Yahoo!?
Get your free -at-yahoo.co.uk address at http://mail.yahoo.co.uk
or your free -at-yahoo.ie address at http://mail.yahoo.ie




From: Petra Wahlbring :      wahlbrin-at-crpgl.lu
Date: Wed, 07 Jun 2000 09:49:03 +0200
Subject: TEM - Searching for Review Article of Biological Applications

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Since I am usually more involved in material analysis I am a little lost on
the biological field. But now I am in the need of a good review article
concerning the application of (analytical) TEM to biological materials. I
am not looking for sophisticated latest developements but rather for basic
applications with some examples, if possible including some examples of
analytical TEM.

Hope to get some input :)

Petra


--------------------------------------------------------------
Dr. Petra Wahlbring
Centre de Recherche Public - Gabriel Lippmann
Laboratoire d'Analyse des Materiaux (LAM)
162a, av. de la Faiencerie L-1511 Luxembourg
tel. +352-466644-402 fax +352-466644-400
e-mail: petra.wahlbring-at-crpgl.lu
Visit our WWW site! http://www.crpgl.lu/~wahlbrin




From: Gary Coulton :      g.coulton-at-ic.ac.uk
Date: Wed, 07 Jun 2000 12:09:24 +0100
Subject: ICHC 2000

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friend

ONLY 8 WEEKS TO GO UNTIL ICHC 2000 (3-8 Sept. York University, UK)

Abstracts are still welcome.

Register by the week or combinations of days.

For more details follow the links to the ICHC 2000 web-site below.

Best wishes

Gary Coulton
Dr. Gary Coulton
Molecular Pathology
Division of Biomedical Sciences
Imperial College School of Medicine
The Sir Alexander Fleming Building
South Kensington
London SW7 2AZ

TEL. ++44 020 7594 3190
FAX. ++44 020 7594 3022
e-mail g.coulton-at-ic.ac.uk


-------------------------------------
Announcing the 11th International Congress of Histochemistry and
Cytochemistry (ICHC 2000)

"Cell Biology and Imaging Tools for the New Century"

September 3-8, 2000, York, United Kingdom

ICHC 2000 comprises 27 symposia addressing latest developments and
applications of histochemistry and cytochemistry in the life sciences
including medicine.

Many leading experts to speak

8 weeks to go! Register for the week or by the day!

For further details go to http://www.med.ic.ac.uk/external/ichc_2000

See you there.




From: jim :      jim-at-proscitech.com.au
Date: Wed, 7 Jun 2000 23:05:26 +1000
Subject: RE: Film vs Digital - the end?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I was happy to leave things at that, but there are a couple of things that Mike
has raised that need an answer:
Your actual quote concerning Ansel Adams was "Having said that and looking at a
print of Ansel Adams, I am glad there is film, though!!" This is rather at odds
with the new proclamation
} " I mentioned Adams because I like his pictures. If he had taken them with
} a digital camera I would have liked them just as much.
Lots of people have made excellent compositions and many have make technically
superior prints. Adam's has excelled by consistently winning on both account.
Its a bit strange, but Mike I'll remind you why you his prints: It is not just
Adam's great compositions but also the incredible sharpness and tonal range,
which is only possible by huge data density and impossible with a 2.5mb
enlarged print. Digital not only has problems with occasional great demands on
print magnification (pixel size limitations) and lack of electron density (too
short exposures), but because TEM negs, like Adam's prints, are capable of a
terrific tonal range and resolution. An image appears better if theoretical
data minima are exceeded.

I have been a stickler for showing possible conflict of interest. I realize now
that I should have made a disclaimer along the way. I did not, simply because
it never occurred to me that I could have a real or imagined gain. My
aspirations are different, but since we are all subject to self-delusion I'll
add the reality:
We have very little mark-up on film and cannot export beyond New Zealand. For
Australian research institutions money has been very tight and equipment sales
have been low for some years. I doubt that at our exchange rate any digital TEM
cameras will be installed here during the next couple of years at least. I
think that my special friend Chuck is more likely to retain more film sales
than ProSciTech may have lost - if anybody was influenced either way.

The word belief is fitting for all of this. Whatever the arguments, I expect
people retain their "film versus digital" beliefs. Mike, we can lead a horse to
water, but just try to make it float on its back. When you do, send a picture
and I won't care if its digital.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Tuesday, June 06, 2000 1:10 AM, Mike Bode [SMTP:mb-at-soft-imaging.com] wrote:
} Jim wrote:
}
} 'I have two replies and Mike my well have four replies to these.'
}
} I'll try. But I agree with you that we should let this thread die. jim,
} it seems you're a bit angy at me. If I unintentionally stepped on your
} toe I apologize. I did not expect that we would agree 100%, as you are
} selling film and I am involved in digital systems. I just hope that some
} other readers found some of this useful. I don't consider myself a
} "digital fanatic". This film vs. digital issue has many more facets,
} some of which we did not even touch, and which can be just as
} entertaining.
}
} Jim wrote:
}
} 'Mike reinforced and strengthened that argument, finishing with the note
} that he is glad for film when looking at prints by Ansel Adams.'
}
} Just for the record: I usually do not take a magnifying glass to photos.
} I mentioned Adams because I like his pictures. If he had taken them with
} a digital camera I would have liked them just as much. I also like some
} modern art paintings. That does not mean we should start drawing what we
} see in the microscopes rather than taking pictures ;-)
}
} Jim wrote:
}
} 'Incidentally, from the outset I cited increased "enlargability" to
} obtain high
} resolution TEM'
}
} Well, you said 'When great enlargements are required film is superior'.
} Perhaps I misunderstood. I thought, the term great enlargements referred
} to the microscope and meant 'small details', which you can take easier
} with a digital camera (no time delay between seeing something and taking
} an image, no mechanical vibration due to film movement, etc.). I have
} said many times before that film may be better if high resolution AND
} large field of view is required.
}
} Jim wrote:
}
} 'in any case: Show Sergey!'
}
} I'm in back-channel correspondence with him.
}
}
}
} Michael
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 1675 Carr St., #105N
} Lakewood, CO 80215
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} From: jim [mailto:jim-at-proscitech.com.au]
} Sent: Sunday, June 04, 2000 2:34 AM
} To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} Subject: RE: Film vs Digital
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I have two replies and Mike my well have four replies to these.
} 1. There has been a parallel discussion concerning resolution of
} film versus
} digital images. Clearly in raw power digital cannot compete since film
} has
} multi gigabyte capacity. I added to that thread that what matters is:
} does
} digital have enough power and that frequently it would. Mike reinforced
} and
} strengthened that argument, finishing with the note that he is glad for
} film
} when looking at prints by Ansel Adams.
} (Ansel Adams until about 30 years ago carted for decades large format
} cameras
} through US National Parks, especially Yosemite, producing fantastic
} landscape
} photographs and books)
} Adams' limited edition prints were contacts of 5x7 and 8x10" inch sheet
} film,
} probably rated at 400 ISO. The line resolution of such prints much
} exceeds our
} eyes' resolution, but still results in superior gradation and detail.
} TEM film,
} even when much enlarged has such details too.
} Why Mike, should we accept 2.5 Mb? That is a splendid file size for SEM,
} and
} because of the limited enlargability of light microscopy (concrete
} ceiling due
} to wavelengths) it's reasonable, but minimal for light microscopy. TEM
} can do
} and deserves better.
}
} Incidentally, from the outset I cited increased "enlargability" to
} obtain high
} resolution TEM, as one of films major advantages, since greater depths
} of field
} at moderate powers makes high powers through photo enlarging a desirable
}
} technique. The small additional magnification yielded by placing a
} digital
} camera lower in the column does not compensate. So greater
} "enlargability" of
} digitals would be desirable, but is limited by pixel size.
}
} 2 The thread was initiated by Sergey. He had problems visualising
} certain
} specimens in dark field. The use of his "beaut" digital TEM camera made
} things
} worse. I pointed out that the shorter exposure reduced the number of
} electrons
} forming the image, hence more noise. I believe that a good part of
} Mike's case
} will be settled in his favour when we hear "Eureka" from Sergey's lab. I
} don't
} doubt that much more can be done with digital now and that further
} improvements
} are on the way. Mike's "solution" may well be possible, but I don't
} believe its
} a snap; in any case: Show Sergey!
} Cheers
} Jim Darley
} ProSciTech Microscopy PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} www.proscitech.com
}
} On Saturday, June 03, 2000 2:58 AM, Mike Bode [SMTP:mb-at-Soft-Imaging.com]
} wrote:
} }
} } Well, let's see:
} }
} } Jim wrote:
} }
} } Mike Bode would use multiple digital
} } exposures. The exposures could be layered and combined into one
} superior
} } image.
} } This image would be made up of more pixel and is formed by more
} } electrons and
} } so would be noise-free and hence could be further enlarged then
} } otherwise
} } possible. Perhaps.
} }
} } No, I did not talk about further enlargements. All I wanted to say is,
} } that a more noise-free image can be achieved by adding multiple
} images,
} } and that this also to some extent helps with drift of the sample
} during
} } acquisition.
} }
} } Jim wrote:
} }
} } How much time is required between exposures to transfer a minimum 10mb
} } image
} } per exposure?
} }
} } How did you arrive at 10 MB? A 1280x1024 image with 16 bit pixel
} } information is about 2.5 MB (uncompressed). We acquire about 10 of
} those
} } per second and transfer them across the PC bus to the display. Putting
} } them on them into Memory might add a few tenth of a second. Writing to
} } HD can be done after all images are acquired.
} }
} } Jim wrote:
} }
} } What would be the total time from focusing to the last exposure?
} } What about Z-drift in the interim requiring objective changes
} }
} } Why would we have to worry about that, if we don't have to worry about
} } that when taking the image on film? In fact, we could take care of
} this
} } by looking at the image between exposures and correct for z-drift.
} } However, as you said, that would add to the overall time and exposure.
} I
} } was comparing a normal dark field image taken on film at 8 seconds
} with
} } acquiring the same image on a "too sensitive" CCD camera by adding up
} 10
} } consecutive .8 second images. Why would the sample drift (in x, y or
} z)
} } substantially more in 8+delta seconds than in 8?
} }
} } Jim wrote:
} }
} } what about the total cost of this additional get-up
} }
} } That of course depends on the microscope and there is no general
} answer.
} } For example on a LEO 912 I believe the blanker is standard. The
} } additional cost to use an acquisition scheme like this with our
} software
} } is $0 plus perhaps a bit of time to write a small macro. On other
} } microscopes one might have to add a beam blanker and perhaps a control
} } mechanism for the beam blanker. But I would guess, that this cost is
} not
} } very high. All modern microscopes are computer controller anyway, so
} it
} } is most likely just a control command that needs to be sent to the
} } microscope over a serial port if the beam blanker is installed. Piece
} of
} } cake.
} }
} } Jim wrote:
} }
} } The mind boggles at a through focus series.
} }
} } You're right here. But I don't think we were talking about
} through-focus
} } series. Incidentally, we do through-focus series on light microscopes
} } and reconstruction routinely. Takes a few images at different focus
} (or
} } for a light microscope: stage) settings. The rest is done off-line.
} } Takes maybe a couple of minutes for about 20 images of about 1kx1k. I
} } agree that TEM is different here and much more complicated due to the
} } complicated Contrast Transfer Function. However, this could in
} principle
} } be sorted out.
} }
} } Jim wrote:
} }
} } Again, I don't doubt that there is now a large place for digital in
} TEM,
} } but
} } its no panacea.
} }
} } I also agree with you on that one. But using the additional computer
} } possibilities of digital imaging might take you further than expected.
} }
} } Michael
} }
} }
} } Michael Bode, Ph.D.
} } Soft Imaging System Corp.
} } 1675 Carr St., #105N
} } Lakewood, CO 80215
} } ===================================
} } phone: (888) FIND SIS
} } (303) 234-9270
} } fax: (303) 234-9271
} } email: mailto:info-at-soft-imaging.com
} } web: http://www.soft-imaging.com
} } ===================================
} }
} }
} }
} } -----Original Message-----
} } } From: jim [mailto:jim-at-proscitech.com.au]
} } Sent: Friday, June 02, 2000 5:15 AM
} } To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} } Cc: 'jim-at-proscitech.com.au'
} } Subject: RE: Film vs Digital
} }
} }
} } The world is full of possible solutions, but are they practical.?
} } To produce high-resolution, dark-field or any others TEM images that
} } require
} } more electrons to form a clear image, Mike Bode would use multiple
} } digital
} } exposures. The exposures could be layered and combined into one
} superior
} } image.
} } This image would be made up of more pixel and is formed by more
} } electrons and
} } so would be noise-free and hence could be further enlarged then
} } otherwise
} } possible. Perhaps.
} } Beam blanking would largely save the specimen from beam damage and
} drift
} } could
} } be compensated for by matching up the digitals. Great.
} } How much time is required between exposures to transfer a minimum 10mb
} } image
} } per exposure? What would be the total time from focusing to the last
} } exposure?
} } What about Z-drift in the interim requiring objective changes and what
} } about
} } the total cost of this additional get-up. The mind boggles at a
} through
} } focus
} } series.
} } When pushing the limits a piece of film seems more effective, cheaper
} } and fa
} } ster.
} } Again, I don't doubt that there is now a large place for digital in
} TEM,
} } but
} } its no panacea.
} } Cheers
} } Jim Darley
} } ProSciTech Microscopy PLUS
} } PO Box 111, Thuringowa QLD 4817 Australia
} } Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} } Great microscopy catalogue, 500 Links, MSDS, User Notes
} } www.proscitech.com
} }
} } On Friday, June 02, 2000 1:24 AM, Mike Bode [SMTP:mb-at-soft-imaging.com]
} } wrote:
} } }
} } } No, it's not really a problem. It's been done with low density
} } } microscopy all the time. Granted, there are some technical aspects
} to
} } be
} } } overcome, but (and I can only speak for ourselves) we have done that
} } on
} } } a number of microscopes. You are of course correct, that 10 images
} at
} } } 0.8 seconds take longer than 8 seconds as the image has to be
} } } transferred, etc. BUT: that's what beam blankers are for. It is
} pretty
} } } straightforward to take an image at 0.8 seconds, then blank the beam
} } } very quickly before taking the next image. That way you get pretty
} } close
} } } to the 8 sec total exposure. If there is no beam blanker on the
} } } microscope, in most cases it can be added.
} } }
} } } I am not sure what you mean by "too sensitive". The cameras are
} } usually
} } } constructed so that 1 electron from the beam creates between a few
} } tenth
} } } to a few counts (these are all statistical data, of course). The
} well
} } } width divided by this sensitivity then determines, how many primary
} } } electrons are needed to fully expose one pixel. For example, if the
} } well
} } } width is 50,000 electrons and the sensitivity is 1 count/electron,
} one
} } } needs 50,000 primary electrons to fill the well. This translates
} into
} } } roughly a 0.4% statistical error.
} } }
} } } } From a practical standpoint: You can take images with most cameras
} } when
} } } the exposure meter on the microscope reads a couple of seconds
} without
} } } overexposing the camera. On the other hand, you can reduce the
} } intensity
} } } of the beam until you see single electron events.
} } }
} } } The one area where CCD cameras may be too sensitive is diffraction.
} } The
} } } normally huge intensity in the transmitted beam often leads to
} } } saturation. In CCDs this can lead to blooming (the intensity spills
} } over
} } } into neighboring pixels). This can be taken care of with special
} chips
} } } that have anti-blooming features, but this usually has some other
} } } drawbacks. Again, this can also be overcome somewhat with multiple
} } } exposures. Film behaves more civilized here, as it simply stops
} } } responding to the electrons, but this makes film more or less
} useless
} } } for quantitative measurements of diffraction patterns. I have done
} } } diffraction with CCDs many times and though it does require some
} } } tweaking, one can get very good results from them.
} } }
} } } Michael Bode, Ph.D.
} } } Soft Imaging System Corp.
} } } 1675 Carr St., #105N
} } } Lakewood, CO 80215
} } } ===================================
} } } phone: (888) FIND SIS
} } } (303) 234-9270
} } } fax: (303) 234-9271
} } } email: mailto:info-at-soft-imaging.com
} } } web: http://www.soft-imaging.com
} } } ===================================
} } }
} } }
} } }
} } } -----Original Message-----
} } } From: jim [mailto:jim-at-proscitech.com.au]
} } } Sent: Wednesday, May 31, 2000 9:37 PM
} } } To: 'Mike Bode'; 'Microscopy-at-MSA.Microscopy.Com'
} } } Subject: RE: Film vs Digital
} } }
} } }
} } }
} }
} ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } America
} } } To Subscribe/Unsubscribe -- Send Email to
} } ListServer-at-MSA.Microscopy.Com
} } } On-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} }
} -----------------------------------------------------------------------.
} } }
} } }
} } } Yes, it could be done "in theory". Somebody would need to figure out
} } the
} } }
} } } software and perhaps modify the hardware. Then we would find that
} the
} } } total
} } } exposure of the specimen to the electron beam maybe a muliple of the
} } } film's
} } } exposure. Afterall, an 8 sec film exposure would not amount in
} digital
} } } to
} } } 10x0.8, but we would require considerable time in between exposures.
} } } Since the
} } } problems in the discussed circumstances are specimen movement and
} beam
} } } damage,
} } } it seems that taking multiple exposures is a poor option.
} } }
} } } Digital cameras are for some situation too sensitive to electron
} } } exposure.
} } } Cutting back on electrons is no option since its the electrons that
} } form
} } } the
} } } image in the first instance.
} } } Much easier in light microscopy . . . insert a neutral density
} } filter.
} } } Cheers
} } } Jim Darley
} } } ProSciTech Microscopy PLUS
} } } PO Box 111, Thuringowa QLD 4817 Australia
} } } Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
} } } Great microscopy catalogue, 500 Links, MSDS, User Notes
} } } www.proscitech.com
} } }





From: Walck, Scott D. :      walck-at-ppg.com
Date: Wed, 7 Jun 2000 09:17:56 -0400
Subject: RE: Ceramic cross-section preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This sounds like a perfect application of the small angle cleavage
technique. You might want to contact Ray Tweston at Univ. of Illinois
because I think that we might have successfully prepared one of these while
I visited there. At any rate, the technique is very inexpensive and you can
prepare sample of superior quality. Get your hands on John McCaffrey and my
paper in the MRS TEM sample Prep IV book (vol 480). It has a detailed
description on how to do it. South Bay Technology sells the Microcleave kit
and you should contact them as well.


-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)



} -----Original Message-----
} From: Ian MacLaren [mailto:maclariz-at-yahoo.co.uk]
} Sent: Wednesday, June 07, 2000 3:28 AM
} To: Microscopy list
} Subject: Ceramic cross-section preparation
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} --------------------------------------------------------------
} ---------.
}
}
} Dear all,
} I'm currently having a few problems preparing good
} cross-section specimens
} of ceramic thin films on LaAlO3 or NdGaO3 substrates.
}
} I can cut them okay, but thinning them often results in
} cracking as the
} thickness goes below 100 microns. I have a Gatan model 623
} disk grinder and
} currently have access to SiC abrasive papers to grit sizes of
} 1500 or 2000,
} and also have the Gatan diamond polishing disks and specimen
} lapping kit.
}
} The problems are as follows:
} If we use the 1500 or 2000 grit SiC, we can often get the
} samples thin
} enough, but with poor surface finish which needs to be improved with
} dimpling (on both sides), we also risk cracking the sample as
} SiC papers
} often contain imperfections.
}
} Using the Gatan specimen lapping kit, I find that as the disk
} grinder is
} altered to grind off a further 10 microns, the edge of the
} sample often
} catches on the diamond disk and tears some of the diamond
} coating off,
} leaving a lump on the surface which then risks catching the
} specimen and
} cracking it.
}
} I have one possible alternative approach which I used in
} Sweden last year
} (with SiAlON ceramics) involving attaching the sample to a
} glass slide and
} polishing it using diamond spray on non-absorbent paper. I
} may consider
} trying this with these new materials.
}
} Other ideas would be welcome, however, as would suggestions of how to
} improve my present method. Please note that I do not have
} the budget to buy
} any expensive new polishing equipment (such as a tripod polisher, for
} instance), so the most welcome suggestions would be those
} that involve
} improvements to current techniques, use of different types of
} polishing
} consumables, etc..
}
} I hope to hear several suggestions, both from users and from
} the companies
} who producing polishing and lapping equipment. Why not post
} them with the
} list so we can all benefit from the sharing of experience?
}
} Best wishes
}
} =====
} Ian MacLaren
} Beijing Laboratory of Electron Microscopy
} Chinese Academy of Sciences, P.O. Box 2724
} 100080 Beijing
} China
} General Email: ian.maclaren-at-physics.org
} Work (esp. large attachments): maclaren-at-image.blem.ac.cn
}
} ____________________________________________________________
} Do You Yahoo!?
} Get your free -at-yahoo.co.uk address at http://mail.yahoo.co.uk
} or your free -at-yahoo.ie address at http://mail.yahoo.ie
}




From: Judith C. Yang :      jyang-at-engrng.pitt.edu (by way of Nestor J.
Date: Wed, 7 Jun 2000 08:34:48 -0500
Subject: post-doc opening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Postdoctoral Position Immediately Available
In situ UHV-TEM of Nanoparticle Reactions and Metal Oxidation

Materials Research Laboratory, University of Illinois at
Urbana-Champaign
and University of Pittsburgh

A postdoctoral position is immediately available in the Materials
Research
Laboratory at the University of Illinois at Urbana-Champaign in the area
of
nano-reactions and in situ UHV-TEM. This position is jointly sponsored
between
Professor Robert Averback (University of Illinois at Urbana-Champaign)
and
Professor Judith Yang (University of Pittsburgh).

The research project is two-fold: 1. Surface oxidation kinetics of
metals
and 2. Nanoparticle reactions/sintering. The research project is to
combine the unique experimental information obtainable from in situ
UHV-TEM
and compare with
theoretical models of these nanoscale reactions. The position requires
a
PhD in physical sciences/engineering. Hands-on experience in TEM
techniques
is highly desirable. The position is open to all qualified candidates
and
has an anticipated duration of 2 years. If you are interested in this
opportunity, please send a resume and names of three references to the
address below.


Dr. Judith C. Yang
Assistant Professor
Dept. of Materials Science & Engineering
848 Benedum Hall
University of Pittsburgh
Pittsburgh, PA 15261

Tel: (412) 624-8613
Fax: (412) 624-8069
jyang-at-engrng.pitt.edu






From: Corneliu Sarbu :      Corneliu.Sarbu-at-mtm.kuleuven.ac.be
Date: Wed, 7 Jun 2000 08:37:34 -0500
Subject: re: TEM - Searching for Review Article of Biological Applicatio

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I can reccomend you the following books:
1. D.C.Joy, A.D.Romig and J.I.Goldstein, "Principles of analytical
electron microscopy", Plenum Press, 1989;
it contains three chapters dedicated to bilogy: chapter 6, 12 and 13
2. J.J.Hren, J.I.Goldstein and D.C.Joy, "Introduction to AEM",
Plenum Press, 1979;
it contains chapters dedicated to biology.
I hope this helps.

Corneliu Sarbu
Dept.MTM
KULeuven
Belgium

- - - - - - - - - - - - - - Original Message - - - - - - - - - - - - - -
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America




Since I am usually more involved in material analysis I am a little lost on
the biological field. But now I am in the need of a good review article
concerning the application of (analytical) TEM to biological materials. I
am not looking for sophisticated latest developements but rather for basic
applications with some examples, if possible including some examples of
analytical TEM.

Hope to get some input :)

Petra


--------------------------------------------------------------
Dr. Petra Wahlbring
Centre de Recherche Public - Gabriel Lippmann
Laboratoire d'Analyse des Materiaux (LAM)
162a, av. de la Faiencerie L-1511 Luxembourg
tel. +352-466644-402 fax +352-466644-400
e-mail: petra.wahlbring-at-crpgl.lu
Visit our WWW site! http://www.crpgl.lu/~wahlbrin


- - - - - - - - - - - - End of Original Message - - - - - - - - - - - -






From: Kim Pierson :      piersokw-at-uwec.edu
Date: Wed, 07 Jun 2000 09:03:26 -0500
Subject: JEOL 2010 LaB6 Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Question:
What is the proper procedure for increasing the filament current on a 2010
with a LaB6?
How fast can you go up?
Should the current be increased at a linear rate or should the rate be
tapered off as you reach saturation?
The manual suggests 30 sec/graduation while the service techs suggest a
much slower rate.

Thanks,
Kim
**************************************************************
Kim W. Pierson, Ph.D.
Dept of Physics & Astronmoy
University of Wisconsin-Eau Claire
(715) 836-5009
FAX 836-3955





From: David Henriks :      Henriks-at-CompuServe.COM
Date: Wed, 7 Jun 2000 10:41:19 -0400
Subject: Ceramic cross-section preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ian:

I have a few comments for you here that may help. This is a little touchy
as I don't want to knock my competitor's equipment. The Model 623 Disk
Grinder that you have is going to cause problems as you describe because of
it's design. The way that grinder works is that you use the dial on the
top to make your sample extend below the base of the grinder. When you
begin polishing, the entire weight of the grinder plus whatever weight you
are applying by hand is being transferred to your very thin 3mm diameter
sample. In addition to the additional weight, you can tend to apply too
much pressure on one side of the sample which would cause the sample to
"catch" on the diamond film as you describe.

We manufacture a series of polishing fixtures that are designed to prevent
these problems. They are gravity feed fixtures. On these fixtures, the
dial gauge at the top simply sets the amount of material that will be fed
into the abrasive film. The only weight your sample sees is the weight of
the central piston (with some of our fixtures, this weight can be
counterbalanced to approach zero). The weight of the rest of the fixture
is never transferred to the sample. Also, the sample surface that is being
polished is always co-planar with the base of the polishing fixture. This
ensures that you cannot apply uneven pressure and that you won't experience
the "catching" that you described.

Now the good news. There is one of these fixtures (Model 150) already in
use in your facility. There is also the Tripod Polisher¨, Model 920
Lapping & Polishing Machine, Model 850 Wire Saw and Model 650 Diamond Wheel
Saw. By separate e-mail,I will give you the details on where these are
located and who you should contact to access them.

I also saw Scott Walck's post on the Microcleave technique. I will include
a PDF of the paper he was talking about by separate e-mail.

If you have any questions or if I can be of any additional assistance,
please contact me directly off-line. I hope this helps.

DISCLAIMER: South Bay Technology produces equipment and supplies as
described above and, therefore, has a vested interest in promoting their
use.

Best regards-

David
Writing at 7:20:30 AM on 06/07/2000

***************************************************************************
************************

David Henriks
Vice President TEL: 800-728-2233
(toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499
San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com

***************************************************************************
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.
Message text written by INTERNET:ian.maclaren-at-physics.org
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Dear all,
I'm currently having a few problems preparing good cross-section specimens
of ceramic thin films on LaAlO3 or NdGaO3 substrates.

I can cut them okay, but thinning them often results in cracking as the
thickness goes below 100 microns. I have a Gatan model 623 disk grinder
and
currently have access to SiC abrasive papers to grit sizes of 1500 or 2000,

and also have the Gatan diamond polishing disks and specimen lapping kit.

The problems are as follows:
If we use the 1500 or 2000 grit SiC, we can often get the samples thin
enough, but with poor surface finish which needs to be improved with
dimpling (on both sides), we also risk cracking the sample as SiC papers
often contain imperfections.

Using the Gatan specimen lapping kit, I find that as the disk grinder is
altered to grind off a further 10 microns, the edge of the sample often
catches on the diamond disk and tears some of the diamond coating off,
leaving a lump on the surface which then risks catching the specimen and
cracking it.

I have one possible alternative approach which I used in Sweden last year
(with SiAlON ceramics) involving attaching the sample to a glass slide and
polishing it using diamond spray on non-absorbent paper. I may consider
trying this with these new materials.

Other ideas would be welcome, however, as would suggestions of how to
improve my present method. Please note that I do not have the budget to
buy
any expensive new polishing equipment (such as a tripod polisher, for
instance), so the most welcome suggestions would be those that involve
improvements to current techniques, use of different types of polishing
consumables, etc..

I hope to hear several suggestions, both from users and from the companies
who producing polishing and lapping equipment. Why not post them with the
list so we can all benefit from the sharing of experience?

Best wishes

=====
Ian MacLaren {




From: Mark West :      mwest-at-ifcsun1.ifisiol.unam.mx
Date: Wed, 7 Jun 2000 09:50:51 -0600 (Hora est‡ndar de MŽxico)
Subject: plant vesicles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I'm trying to do immunogold studies on a membrane protein in a preparation
of plant vesicles (from minus 80ĽC storage). I did an Epon embedding with
standard procedure (glut and osmium) to check the ultrastructure and got
decent-looking vesicles, but side by side an LR White prep (0.5% glut, 2%
paraformaldehyde, without osmium) looks like a disaster, with granular
material, vague membranish-like structures and whorls of membranes. Does
anyone have any tips for LR White preps of vesicles, or any other ideas
for embedding (resins, fixing, dehydration, embedding) to favour the
immunogold process.

Thanks,

Mark






********************************************
Mark West,
Unidad de Microscopia Electronica,
(Electron Microscopy Unit)
Instituto de Fisiologia Celular,
Universidad Nacional Autonoma de Mexico,
04510 Mexico D.F.

tel (unidad/lab) *(525) 622 5610*
(casa/home) (525) 619 3020
Fax (525) 616 2282
********************************************





From: Michelle.Taurino-at-aventis.com
Date: Wed, 7 Jun 2000 11:59:26 -0500
Subject: TEM - Searching for Review Article of Biological Applications

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Petra-

A book that I have found to be very resourceful is
Biomedical Electron Microscopy by Arvid B. Maunsbach and Bjorn Afzelius.

Hope this is useful to you as well.
Regards-
Michelle Taurino
Aventis Pharmaceuticals
Bioimaging and Molecular Histology
Tel-908-231-3357
Fax-908-231-3962
e-mail: Michelle.Taurino-at-aventis.com


-----Original Message-----
} From: Petra Wahlbring [mailto:wahlbrin-at-crpgl.lu]
Sent: Wednesday, June 07, 2000 3:49 AM
To: microscopy-at-sparc5.microscopy.com




Since I am usually more involved in material analysis I am a little lost on
the biological field. But now I am in the need of a good review article
concerning the application of (analytical) TEM to biological materials. I
am not looking for sophisticated latest developements but rather for basic
applications with some examples, if possible including some examples of
analytical TEM.

Hope to get some input :)

Petra


--------------------------------------------------------------
Dr. Petra Wahlbring
Centre de Recherche Public - Gabriel Lippmann
Laboratoire d'Analyse des Materiaux (LAM)
162a, av. de la Faiencerie L-1511 Luxembourg
tel. +352-466644-402 fax +352-466644-400
e-mail: petra.wahlbring-at-crpgl.lu
Visit our WWW site! http://www.crpgl.lu/~wahlbrin




From: Caroline Schooley :      schooley-at-mcn.org
Date: Wed, 7 Jun 2000 10:01:04 -0700
Subject: Re: SEM: agency wants SEM images?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} My SEM facility has just been approached, and in the interest in
} providing what they want for a price based on an average, can I query
} the SEM community as to what they charge for commercial release of
} their SEM imagery. I'm looking for $/image, but if you charge on any
} type of sliding scale, I'd be interested in that too. I believe as
} well I shouldn't be undercutting commercial facilities, so I'm
} especially interested in these numbers.
} Reply to me direct, and I'll respond back to the list with the
} average, max and min, and no names.

shAf-

Please start with an inquiry to your university administration; you might
save yourself a lot of grief. When I was running an interdepartmental lab
at U.C. Berkeley, there were very specific university-wide rules that I was
required to follow for such usage.

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: Marti, Jordi :      jordi.marti-at-honeywell.com
Date: Wed, 7 Jun 2000 13:02:55 -0700
Subject: RE: JEOL 2010 LaB6 Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Kim:

We have a 2010 and our procedure is to bring the filament dial (labeled
off-10) to position #3. Then we go for coffee. After 20 min. or so we raise
it to # 4, after a couple of minutes to #5 etc. until we get to the stop (
in our case we have it at about 7 1/2). Even at this point we have filament
drift for quite some time, so , we are not ready to take pictures for
another 1/2 hr or so. By the way, we have the bias settings at 7 and 7 which
is high. I would be interested in knowing what your settings are.

I hope this helps

Jordi Marti





From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Wed, 7 Jun 2000 20:48:55 +0100
Subject: Re: TEM:beam damage vs voltage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 14:19 +0200 3/6/00, Jonathan Barnard wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

My understanding is as follows ....

There are different damage mechnisms at different voltages.

At lower voltages ( {200 kV) the main damage mechanism is, in effect,
specimen heating. And, the lower the voltage, the greater the
interaction (elastic scattering) with the specimen - so more damage
but greater contrast, the main reason users looking at resin sections
prefer lower voltages.

Increasing the voltage reduces specimen damage and tends to improve
resolution but at the expense of contrast. As the interest in
biological TEM moves to molecular biology, resolution is more
important. But, at high resolution, phase contrast becomes the
dominant factor in producing image constrast and phase constrast can
be significantly improved - without loss of resolution - by using
highly coherent FEG sources.

At voltages between 300kV and 400 kV, the energy of the electron
becomes sufficient to cause "direct" radiation damage - atoms are
displaced. At this point, specimen damage rates increase. The
particular voltage depends on the amtomic number of your specimen.

Damage rates can also be reduced by cooling. in particular, cooling a
specimen to liquid helium temperatures allows significantly greater
electron exposure before damage occurs.

So, for optimum imaging of molecular specimens you need 300 kV and a
FEG gun plus a liquid helium stage.

Regards,
--
Larry Stoter
JEOL (UK) Ltd
Silver Court, Watchmead, Welwyn Garden City, AL7 1LT, United Kingdom
tel: +44-(0)1707-377117, fax: +44-(0)1707-373254, e-mail: larrys-at-jeoleuro.com




From: Dusevich, Vladimir :      DusevichV-at-umkc.edu
Date: Wed, 7 Jun 2000 17:11:48 -0500
Subject: RE: NEED HELP WITH EDS ON A JEOL 100S

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For EDS checking you can use specimen
from Cu or Al foil so that it
will have any angle, including 30-60
degrees. But I am afraid that if there
is no spectra at all with 10 degree
tilt then a problem is more complicated.

Vladimir

Vladimir M. Dusevich, Ph.D.
Electron Microscope Lab Manager
3127 School of Dentistry
650 E. 25th Street
Kansas City, MO 64108-2784

Phone: (816) 235-2072
Fax: (816) 235-5524
Web: http://www.umkc.edu/dentistry/microscopy

} -----Original Message-----
} From: "MGMANDERS-at-aol.com"-at-sparc5.microscopy.com
} [mailto:"MGMANDERS-at-aol.com"-at-sparc5.microscopy.com]
} Sent: Sunday, June 04, 2000 10:41 PM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: NEED HELP WITH EDS ON A JEOL 100S
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListser} ver/FAQ.html
}
}
} --------------------------------------------------------------
} ---------.
}
}
} A fellow colleague has bought a JEOL 100s and wants and
} needs EDS X-ray
} Analysis. He's mounted his horizintal EDS detector , but can
} not get sample
} X-rays from the speciman. Contacting JEOL he found the
} problem to be a tilt
} problem. The 10 degree tilt from the normal SEG only tilts
} 10 degrees and a
} 30 to 60 degree tilt is necessary. At one time I was told
} that special
} sample holders and or double gap pole pieces were availabe.
} He wishs to find
} any one with a 100s for parts needed or information which
} would allow X-ray
} analysis. Please reply to mgmanders-at-aol.com or the list server.
}
} Mike Manders
}
}
}




From: Rick Powell at Nanoprobes :      rpowell-at-nanoprobes.com
Date: Wed, 7 Jun 2000 21:49:08 -0500
Subject: TEM: Set-up, operating and facilities costs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello List Members:

We have the opportunity to acquire a TEM, but budget constraints mean that
have to consider the cost of acquiring and operating this instrument
carefully against alternatives. We plan to use it principally to study the
morphology, behavior and size distribution of colloidal gold particles and
other types of metal nanoparticles.

Can anyone give any information on what the maintenance, operating and
facilities requirements and costs would be for this instrument, or point me
to a good source? What renovations would be necessary to accommodate it -
darkroom, ancillary equipment, cooling, water and power supplies, darkroom
facilities? How much should we budget for supplies and consumables? Since
we have no-one currently trained to use it, how long would it take to
learn, and how much maintenance (time, supplies, contracts) would it
require?

Thanks in advance,

Rick Powell


**********************************************************************
* - PLEASE NOTE OUR NEW ADDRESS - EFFECTIVE FEBRUARY 15, 2000 - *
* *
* NANOPROBES, Incorporated | US Toll-free: (877) 447-6266 *
* 95 Horse Block Road | Tel: (919) 510-0590 *
* Yaphank, NY 11980-9710, | Fax: (919) 510-0590 *
* USA | rpowell-at-nanoprobes.com *
* *
* NOW EASY TO FIND ON THE WEB: http://www.nanoprobes.com *
**********************************************************************






From: Barbara Foster :      mme-at-map.com
Date: Wed, 07 Jun 2000 23:34:37 -0400
Subject: Re: UV optics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Don,

Both Zeiss and Leica have long-standing reputations for quartz optics with
transmission into the UV (typically, down to about 220nm). You are going to
need more than just the objective; the binoc also needs to have a UV
transmitting prism. A tip: inquire about microscopes used for
microspectrophotometry or in semiconductor applications. From the
biological/biomed realm: microscopes used for Fura or Indo studies.

Try Tom Calahan at Zeiss (914-681-7733) and Jan Hinsch at Leica/Allentown
(201-236-5905).

Nikon has also been agressive in lens development, but I haven't had direct
experience with any UV optics. Call Stan Schwartz (516-547-8529). ... and
at Olympus: Reinhard Enders (516-844-5000).

Good hunting!
Barbara Foster
Microscopy/Microscopy Education
125 Paridon Street, Suitee 102
Springfield, MA 01118

PH: 413-746-6931 FX: 413-746-9311 email: mme-at-map.com

Customized, on-site short courses in all areas of microscopy, sample prep,
and image analysis.
"Why didn't they teach us that sooner?" Probably because no one called MME!






At 03:27 PM 6/6/00 -0400, donald j marshall wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From: Barbara Foster :      mme-at-map.com
Date: Thu, 08 Jun 2000 00:12:10 -0400
Subject: Re: digital OMs for an EM lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Edorado,


Suggest you buy a good upright stand which you can upgrade as your needs
develop. I strongly suggest a combination of both transmitted and
reflected light options.


Your feelings re: investing in the microscope vs. a high end camera at
this stage are well founded.


Re: contrast techniques... I think that you might find Hoffman Modulation
Contrast a better bet for what you are doing than Phase. Hoffman can be
used singly for looking at surfaces (ex: your coatings) or in combination
with polarized light. It is a good complement to DIC, which won't work if
your powders, etc. respond to Pol.


All of the "big 4" (Olympus, Nikon, Leica, and Zeiss) have good
equipment. The issues I would add to your shopping list might include
how comfortable you feel operating a specific microscope (like trying on
a coat or test-driving a new car) and how supportive your local dealer or
representative is.


Your comment re: EM was interesting.... I have been doing a lot of work
over the past 6 months with commercial companies who look at many of the
applications you cited. Universally, they were astounded at the
information they could get from the Light Microscope, quickly and with
minimal sample prep. I am currently on assignment teaching a group of
very competent EM people about Light Microscopy... and they are equally
amazed. All by themselves, they came to the conclusion that they should
take a look first with the Light Microscope and then, if necessary, go
the SEM. (I added that they should also take a quick look with the stereo
first).


By the way, you may be interesting in "Optimizing Light Microscopy", both
as a reference for your lab and a text for your students. Details are on
our website.


Happy shopping!

Barbara Foster

Microscopy/Microscopy Education

125 Paridon Street, Suitee 102

Springfield, MA 01118


PH: 413-746-6931 FX: 413-746-9311 email: mme-at-map.com

www.MME-Microscopy.com/education


Customized, on-site short courses in all areas of microscopy, sample
prep, and image analysis.

"Why didn't they teach us that sooner?" Probably because no one called
MME!



At 09:47 AM 6/7/00 +0200, Bemporad, Edoardo wrote:

} } } }

{excerpt}

{fontfamily} {param} Arial {/param} {smaller} I am going to buy one ore two OM
for our EM lab (XL30 and CM120), trying to convince myself that one brand
is better than the other (quality/price rate included in the
evaluation!).

{/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} In our lab we do mainly
metallographic analysis, interface studies of wear resistant coatings,
and catalytic powders characterization, but I guess that the OM will be
used for a wider range of investigation (W/O and O/W emulsions, asbestos,
..) and for didactical scopes too.

{/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} So EPI and DIA illuminations,
BF, DF, NIC, phase contrast, pol, I don't think we will need
fluorescence; forgotten something?

{/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} What about an inverse
microscope? or always better two standard ones? {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} I have about 35K$ budget to make
everybody happy (research group and students), and my position is to
prefer spending on the microscope (or microscopes) rather than on a
high-level digital camera. I read some threads about it here and I think
something like a Nikon coolpix 990 (if it will works! :-) ) will be
enough. (other opinions or point of views?)

{/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} I do not have a so deep
experience in OM but I was wandering if there are some key feature to
keep in mind for an equipment that will be used in a EM lab, considering
that where I will not able to go with the OM (depth of field, resolution)
I can use our EM!

{/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} Anybody have suggestion (base
set, optional...)? {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} I will post a report with the
collected hints; Thank You . {/smaller} {/fontfamily}


{fontfamily} {param} Arial {/param} {smaller} Dr. Eng. Edoardo Bemporad, Ph.
D. {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} Assistant Professor of Materials
Science {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} University of Rome "Roma Tre"
(Italy) {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} Dipartimento di Ingegneria
Meccanica e Industriale {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} (Department of Mechanical and
Industrial Engineering) {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} Via Vasca Navale 79 - 00146
Rome, Italy {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} Tel:
+39 {/smaller} {/fontfamily} {fontfamily} {param} Arial {/param} {smaller} 06
5517.3293 {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} Fax:
+39 {/smaller} {/fontfamily} {fontfamily} {param} Arial {/param} {smaller} 06
5517.3256 {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} LIME Lab Tel:
+39 {/smaller} {/fontfamily} {fontfamily} {param} Arial {/param} {smaller} 06
5517.3200 {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} LIME Web Site:
http//materials.dimi.uniroma3.it/lime {/smaller} {/fontfamily}

{fontfamily} {param} Arial {/param} {smaller} E-Mail:bemporad-at-uniroma3.it
{/smaller} {/fontfamily}


{/excerpt} { { { { { { { {




{/x-rich}



From: Barbara Foster :      mme-at-map.com
Date: Wed, 07 Jun 2000 23:47:15 -0400
Subject: Re: Help? Cleaning Optical Lenses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Gina,

Most of the lenses should come off, either directly or with the help of a
small Allen wrench.

My favorite approach to lens cleaning: "Puff, Huff, and Swirl"....
1. Puff off any debris or rough material which might scratch the lens using
either a puffer (available from photographic supply houses) or just
"puffing" the dry air from your mouth with your cheeks (easier to demo than
explain).
2. Huff on the lens, using the warm, moist air from deep in your lungs, to
deposit a fog of moisture on the lens then
3. Quickly and gently, remove the fog with a clean Q-tip (100% cotton ) or
lens tissue (NOT Kim-Wipe!), starting at the middle of the lens and
continuing in a spiral to the outer edge. Do not use the same area on the
swab or tissue more than once; they will collect debris which can scratch
the delicate coating.
4. If the dirt is persistent (ex: mascara, fingerprint, oily residue), dip
the tip of a cotton swab in a good lens cleaning solution (also available
at photo supply houses), shake off any excess, then repeat the "swirl" step.

I recently had a client who had a neat moist towelette made by Uvex. It
worked really well, even on oil. I will have to find the contact info, so
if you are interested, please email me off line.

Best regards,
Barbara Foster
Microscopy/Microscopy Education
125 Paridon Street, Suitee 102
Springfield, MA 01118

PH: 413-746-6931 FX: 413-746-9311 email: mme-at-map.com

Customized, on-site short courses in all areas of microscopy, sample prep,
and image analysis.
"Why didn't they teach us that sooner?" Probably because no one called MME!


At 11:43 PM 6/6/00 -0500, wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From: Ron Doole :      ron.doole-at-materials.oxford.ac.uk
Date: Thu, 8 Jun 2000 08:37:05 +0100 (BST)
Subject: Re: JEOL 2010 LaB6 Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi Kim,

The real question is why should you increase the LaB6 heating current
slowly? As I understand it there are two reasons. Firstly if the tip is
new, the gun has been serviced or the gun has not been run at 200kV for
some time it is important to run up slowly to ensure that you do not get
any outgassing from the tip or other gun components that might cause a
flashover and possible damage. Secondly if the machine is regularly used
at 200kV and the gun is is good condition (normal case) then you want to
avoid any thermal shock to the LaB6 tip which could cause damage or
misalignment. The same applies to cooling the tip down after use.

In the first case I would certainly take several 10s of minutes while
carefully watching the vacuum gauge and the emission current meter (and HT
stability if you are connected to an oscilloscope). Exact time would
depend on the condition but maybe 30 minutes to heat a new tip and 2
minutes for a normal tip in good condition. I would take a minute to cool
down a tip.

In both cases the greatest heating effect is a square of the control
position (power is proportional to V^2 or I^2) and this should be taken
into acount. I heat to about 3.5 in 25 to 30 seconds then decrease my
heating speed as I approach the stop (set at 5 in my case, bias 5.5). The
particular setting will depend on the type of LaB6 tip used, the wenhelt
to tip distance and how you run the emission. We use 5uA emission with the
tip slightly undersaturated as it seems to give good results and
reasonable long life at the high mags that we use. For the smallest probes
we may desaturate further to reduce the probe size.

I am aware from visitors that we have from other sites that this is
quicker than many people but it seems impractical to spend 30 minutes to
run up the tip every time you want to change a specimen when we want to
look at several in a session. Our tip life des not seem to be any worse
than those who take much longer.

Regards,
Ron



} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Question:
} What is the proper procedure for increasing the filament current on a 2010
} with a LaB6?
} How fast can you go up?
} Should the current be increased at a linear rate or should the rate be
} tapered off as you reach saturation?
} The manual suggests 30 sec/graduation while the service techs suggest a
} much slower rate.
}
} Thanks,
} Kim
} **************************************************************
} Kim W. Pierson, Ph.D.
} Dept of Physics & Astronmoy
} University of Wisconsin-Eau Claire
} (715) 836-5009
} FAX 836-3955
}
}
}

===========================================================================
Mr. Ron Doole e-mail ron.doole-at-materials.ox.ac.uk
Department of Materials, phone +44 (0) 1865 273701
University of Oxford, fax +44 (0) 1865 283333
Parks Road.
Oxford. OX1 3PH. UK.
============================================================================





From: jim :      jim-at-proscitech.com.au
Date: Thu, 8 Jun 2000 11:09:37 +1000
Subject: RE: plant vesicles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


LR White won't look as good as Spurr's, but membrane whorls are indicative of
insufficient fixation. Os is the better lipid (therefore membrane) fixative,
since you cannot use Os, you may need to increase the time and or concentration
of GA.
Cheers
Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Thursday, June 08, 2000 1:51 AM, Mark West
[SMTP:mwest-at-ifcsun1.ifisiol.unam.mx] wrote:

}
} Hi,
}
} I'm trying to do immunogold studies on a membrane protein in a preparation
} of plant vesicles (from minus 80oC storage). I did an Epon embedding with
} standard procedure (glut and osmium) to check the ultrastructure and got
} decent-looking vesicles, but side by side an LR White prep (0.5% glut, 2%
} paraformaldehyde, without osmium) looks like a disaster, with granular
} material, vague membranish-like structures and whorls of membranes. Does
} anyone have any tips for LR White preps of vesicles, or any other ideas
} for embedding (resins, fixing, dehydration, embedding) to favour the
} immunogold process.
}
} Thanks,
}
} Mark
}
}
}
}
}
}
} ********************************************
} Mark West,
} Unidad de Microscopia Electronica,
} (Electron Microscopy Unit)
} Instituto de Fisiologia Celular,
} Universidad Nacional Autonoma de Mexico,
} 04510 Mexico D.F.
}
} tel (unidad/lab) *(525) 622 5610*
} (casa/home) (525) 619 3020
} Fax (525) 616 2282
} ********************************************
}





From: Per Hšrstedt :      per.horstedt-at-pathol.umu.se
Date: Thu, 08 Jun 2000 13:44:10 +0200
Subject: Re: Still alive?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Kris,
same thing happened to me, but re-subscribing took care of the problem.
Mailed to listserver, but no respons at all.


} Dear Microscopists,
} I'm just curious if something is wrong on my end or I'm unsuscribed against
} my will. I did not receive any posting during the past two-three weeks, and
} this is impossible in view of previously received 10-20 postings/day.
} Is there any explanation?
} Kris
} Dr. Kristof Kovacs
} Associate Professor
} President, Hungarian Society for Microscopy
} Phone: +36-(88)-421-684
} Fax: +36-(88)-328-643
} Mailing Address:
} University of Veszprem, P.O.Box 158, Veszprem
} H-8201 Hungary

Yours sincerely

Per Hšrstedt
Department of Medical Biosciences
Pathology
Unit for Electron Microscopy
University of UmeŚ
S-90187 UmeŚ
Sweden

phone int-46-90-7851541
fax int-46-90-7851215





From: prenitzer,brenda s :      prenitzer-at-lucent.com
Date: Thu, 8 Jun 2000 08:54:59 -0400
Subject: Re: TEM:beam damage vs voltage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Literature indicates that specimen damage due to heating decreases with
increasing accelerating voltage; however, there is a trade off because at
higher voltages materials fall victim to knock-on and sputtering damage. A
very good overview (with references) of specimen/beam interactions and beam
damage can be found on PP 49-55 "Transmission Electron Microscopy" by
Williams and Carter 1996 Plenum Press ISBN: 0-306-45342-X

Brenda I. Prenitzer, Ph.D.
Member of Technical Staff
Cirent Semiconductor (Lucent Technologies)
9333 S. John Young Parkway
6D-Lab
Orlando, FL 32819-8612

Phone: 407 371 7108
Fax: 407 371 6999

prenitzer-at-lucent.com
bsp101-at-worldnet.att.net


-----Original Message-----
} From: Larry Stoter [mailto:LPS-at-teknesis.demon.co.uk]
Sent: Wednesday, June 07, 2000 3:49 PM
To: Jonathan Barnard; microscopy-at-sparc5.microscopy.com


At 14:19 +0200 3/6/00, Jonathan Barnard wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

My understanding is as follows ....

There are different damage mechnisms at different voltages.

At lower voltages ( {200 kV) the main damage mechanism is, in effect,
specimen heating. And, the lower the voltage, the greater the
interaction (elastic scattering) with the specimen - so more damage
but greater contrast, the main reason users looking at resin sections
prefer lower voltages.

Increasing the voltage reduces specimen damage and tends to improve
resolution but at the expense of contrast. As the interest in
biological TEM moves to molecular biology, resolution is more
important. But, at high resolution, phase contrast becomes the
dominant factor in producing image constrast and phase constrast can
be significantly improved - without loss of resolution - by using
highly coherent FEG sources.

At voltages between 300kV and 400 kV, the energy of the electron
becomes sufficient to cause "direct" radiation damage - atoms are
displaced. At this point, specimen damage rates increase. The
particular voltage depends on the amtomic number of your specimen.

Damage rates can also be reduced by cooling. in particular, cooling a
specimen to liquid helium temperatures allows significantly greater
electron exposure before damage occurs.

So, for optimum imaging of molecular specimens you need 300 kV and a
FEG gun plus a liquid helium stage.

Regards,
--
Larry Stoter
JEOL (UK) Ltd
Silver Court, Watchmead, Welwyn Garden City, AL7 1LT, United Kingdom
tel: +44-(0)1707-377117, fax: +44-(0)1707-373254, e-mail:
larrys-at-jeoleuro.com




From: Peter Bond :      P.Bond-at-plymouth.ac.uk
Date: Thu, 8 Jun 2000 08:31:38 -0500
Subject: LRWhite flat embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all

I remember a message some time ago, I think, suggesting a way to flat
embed in silicon molds using LRWhite resin. How the top was sealed
escapes me. Could anyone with a better filing system and/or memory
please send me a message or a reference on this.

Thanks.

Pete
--
Peter Bond
Plymouth Electron Microscopy Unit
University of Plymouth
Drake Circus
Plymouth
Devon
UK PL4 8AA
Tel/Fax: 01752 233092
email: pbond-at-plymouth.ac.uk






From: Ken Tiekotter :      tiekotte-at-up.edu
Date: Thu, 8 Jun 2000 08:35:01 -0500
Subject: SEM/TEM available...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

If you are looking for an inexpensive TEM and/or SEM, The University of
Portland has two such units you might be interested in.

1) Zeiss EM9s2, transmission electron microscopy, with additional high
voltage tank, replacement for vacuum tubes, fuses, new filaments, lots of
film cassettes, and film, original schematics, and operation manuals.

2) ETEC AutoScan, scanning electron microscope, with 60 and 90 degree
stages, several reconditioned column liner tubes, new apertures, new YAG
scintillator in original container, original schematics and operating
manuals, plus, video tapes on operation and use. This instrument was
original manufactured for INTEL and has a 50kV power supply.

Both units are operational!

Any interested parties or individuals can contact me off the server.

Ken Tiekotter, Adjunct Professor
The University of Portland
Department of Biology
5000 N. Willamette Blvd.
Portland, OR 97203 USA

(503) 413-5391







From: Siegel, Stephen :      ssiegel-at-EMSL.com
Date: Thu, 8 Jun 2000 09:56:58 -0400
Subject: TEM Analyst Needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


EMSL Analytical Inc. (Westmont, NJ) is interested in hiring a Transmission
Electron Microscopy Analyst (TEM) for its NJ Corporate Office. EMSL is the
world leader in asbestos analysis since 1981 with over 20 laboratory
locations worldwide.

Responsibilities include the preparation and analysis of air, bulk, and
water samples submitted by clients for asbestos content and other specialty
asbestos projects. College Degree in Materials Science or related field and
past experience with electron microscopy analysis preferred. . The ideal
candidate would be a detail orientated individual able to follow lab
protocols and procedures and able to work in a fast paced environment.

Full benefits package with salary commensurate with experience. Interested
individuals send resume and salary requirements to Stephen Siegel, CIH
(ssiegel-at-emsl.com) or fax to 856-858-4960.

Stephen Siegel
Stephen Siegel, CIH
Asbestos Lab Manager
EMSL Analytical, INC
107 Haddon Avenue
Westmont, NJ 08108
Phone:800-220-3675x1209 Fax:609-858-4960
email:ssiegel-at-emsl.com




From: garyliechty-at-att.net
Date: Thu, 8 Jun 2000 08:50:50 -0500
Subject: Re: Ceramic cross-section preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ian,

Your main problem is with the use of SiC for grinding
and thinning.

Silicon carbide paper should not be used to prepare any
ceramic material, it is not hard enough to cut it
efficiently without damaging the substructure. SiC
becomes dull quite rapidly, with ceramic a dull abrasive
actually cracks the structure and causes pullout and
excessive chipping. You are compounding your problem
with the weight of the tool being used as well. While
the disc grinder is a well designed product and very
useful, it is quite heavy for samples of certain
thicknesses not to mention the additional pressure you
are applying is not helping any.

The solution is to use Diamond Lapping Film and apply
some Diamond Extender at the final stages of thinning to
reduce the surface tension between the water and the
sample. This reduces the sheer stress being applied to
the sample and will drastically reduce the possibility
of cracking.

Should you have interest, we offer a polishing machine
"The MultiPrep System" that allows you to prepare the
sample without the possibility of the tool being
misaligned or mishandled (tilted on edge) during prep.
Only the sample makes contact with the abrasive and the
plane of polish remains in tact throughout the
polishing/grinding process. A dial indicator (1 micron
increments) allows you to preset a known amount of
material to be removed and there is no operator
intervention until the sample is done. The amount of
force applied to the sample is constant regardless of
the operator.

If you wish to get more information about the MultiPrep
System, you may visit our website:
www.alliedhightech.com or contact me in person at 310-
635-2466.

Sincerely,

Gary Liechty
Manager, Technical Products
Allied High Tech Products, Inc.
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear all,
} I'm currently having a few problems preparing good cross-section specimens
} of ceramic thin films on LaAlO3 or NdGaO3 substrates.
}
} I can cut them okay, but thinning them often results in cracking as the
} thickness goes below 100 microns. I have a Gatan model 623 disk grinder and
} currently have access to SiC abrasive papers to grit sizes of 1500 or 2000,
} and also have the Gatan diamond polishing disks and specimen lapping kit.
}
} The problems are as follows:
} If we use the 1500 or 2000 grit SiC, we can often get the samples thin
} enough, but with poor surface finish which needs to be improved with
} dimpling (on both sides), we also risk cracking the sample as SiC papers
} often contain imperfections.
}
} Using the Gatan specimen lapping kit, I find that as the disk grinder is
} altered to grind off a further 10 microns, the edge of the sample often
} catches on the diamond disk and tears some of the diamond coating off,
} leaving a lump on the surface which then risks catching the specimen and
} cracking it.
}
} I have one possible alternative approach which I used in Sweden last year
} (with SiAlON ceramics) involving attaching the sample to a glass slide and
} polishing it using diamond spray on non-absorbent paper. I may consider
} trying this with these new materials.
}
} Other ideas would be welcome, however, as would suggestions of how to
} improve my present method. Please note that I do not have the budget to buy
} any expensive new polishing equipment (such as a tripod polisher, for
} instance), so the most welcome suggestions would be those that involve
} improvements to current techniques, use of different types of polishing
} consumables, etc..
}
} I hope to hear several suggestions, both from users and from the companies
} who producing polishing and lapping equipment. Why not post them with the
} list so we can all benefit from the sharing of experience?
}
} Best wishes
}
} =====
} Ian MacLaren
} Beijing Laboratory of Electron Microscopy
} Chinese Academy of Sciences, P.O. Box 2724
} 100080 Beijing
} China
} General Email: ian.maclaren-at-physics.org
} Work (esp. large attachments): maclaren-at-image.blem.ac.cn
}
} ____________________________________________________________
} Do You Yahoo!?
} Get your free -at-yahoo.co.uk address at http://mail.yahoo.co.uk
} or your free -at-yahoo.ie address at http://mail.yahoo.ie
}






From: Arey, Bruce W :      bruce.arey-at-pnl.gov
Date: Thu, 08 Jun 2000 07:24:57 -0700
Subject: JEOL 840 SEM Vacuum Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have JEOL 840 SEM that has develop a leak at the specimen chamber door
interface and will not hold a vacuum. We have changed the o-ring around the
door, look for nicks, scratches any other damage around the door interface, none
found. We have also check the roughing valve (V4, LV3 and the pressure relief
valve) for leaks. Pressure holds in the gun area. The large door is only opened
when we have to replace or add a new attachment to the chamber and thats about
once a year. We have a vacuum specimen exchange port to enter and retrieve
samples. We have also check that seal and it also checks out OK.
We have had 2 JEOL engineers take a look at the system and at this point no luck
finding the cause or solution. You can respond to me off-line at
bruce.arey-at-pnl.gov
Thanks

Bruce W. Arey

PNNL-Battelle
Associate Scientist
Microstructural Characterization
bruce.arey-at-pnl.gov
509-376-3363
fax 509-376-6308






From: Yves Giroux :      ygiroux-at-istar.ca
Date: Thu, 08 Jun 2000 10:34:27 -0400
Subject: PGT parts for anyone

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hi:

I am about to tear the guts out of an old PGT System4 console. I need some
of the electronics, but will toss the rest unless someone would like it.

Jonathan Krupp
Microscopy & Imaging Lab
University of California
Santa Cruz, CA 95064
(831) 459-2477
jmkrupp-at-cats.ucsc.edu












From: Bruce Brinson :      brinson-at-cnst.rice.edu
Date: Thu, 08 Jun 2000 09:47:26 -0500
Subject: Re: JEOL 2010 LaB6 Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


hello Kim, Jordi & interested parties,
We have a filament ramp circuit on our LaB6 2010 that controls the filament
current. It is an accessory offered by Jeol. Jeol has set it to take about 8
minutes to bring the filament to max current. The max. current is still
controlled by the (preset) filament current knob on L1. Once the filament is
hot, when we change samples or other wise have the filament current off for
{5mi. we use the "quick timer" feature of this circuit which brings the filament
up in ~90 seconds.
The max. filament current is set to run just under saturation (so you can
just barely see the X when the beam is converged). The bias control is used to
set beam current to 10 uA. The setting changes with KV & filament wear. The
initial bias setting is a function of the physical position of the filament
relative to the Wehnelt cap. It changes every time the filament or cap are
serviced. As the tip wears the difference in bias settings say between 100 &
200 KV increases & both move up.
We are a multiple user facility. Our filament sees a of cycles & we always
have novice users. Filament life times are 400-1000 hours. This tends to track
with the novice user density.
As far as the drift Jordi mentioned. This is a non issue here & may be in
the filament design. We use the Gimble Phillips 60-06. I do a lot of carbon
work & can pretty much nail 3.4A when I get a beam. Yes the images are sharper
after we have been running a bit.

Bruce Brinson
Rice U.

usual disclaimer... no financial interest in companies mentioned.

Marti, Jordi wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Kim:
}
} We have a 2010 and our procedure is to bring the filament dial (labeled
} off-10) to position #3. Then we go for coffee. After 20 min. or so we raise
} it to # 4, after a couple of minutes to #5 etc. until we get to the stop (
} in our case we have it at about 7 1/2). Even at this point we have filament
} drift for quite some time, so , we are not ready to take pictures for
} another 1/2 hr or so. By the way, we have the bias settings at 7 and 7 which
} is high. I would be interested in knowing what your settings are.
}
} I hope this helps
}
} Jordi Marti
}





From: John J. Bozzola :      bozzola-at-siu.edu
Date: Thu, 8 Jun 2000 11:13:17 -0500
Subject: Stereometry by computer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A user of our facility is interested in making height measurements
from specimens viewed in the SEM. I am aware of the conventional way
of doing this: stereo pairs and optical viewer (with stereometer
parallax corrections). Is there a more modern (computerized) way of
doing this, say with anaglyphs?

Thank you.

John B.


####################################################################
John J. Bozzola, Ph.D., Director
Micro-Imaging and Analysis Center
750 Communications Drive - MC 4402
Southern Illinois University
Carbondale, IL 62901 U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################




From: Al :      alchazz-at-ix.netcom.com
Date: Thu, 8 Jun 2000 07:59:26 -0500
Subject: Re: SEM: agency wants SEM images?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


I work for an analytical lab, Analytical Answers, Inc., located in Woburn, MA.

We charge $200/hr for SEM imaging, high resolution SEM imaging is more. The
customers are free to do what they want with the SEM images.

They paid for them, they own them!

Al Jaszek




-------------------------------------
*Causa latet, eventus est notissimus*
-------------------------------------






From: DrJohnRuss-at-aol.com
Date: Thu, 8 Jun 2000 13:40:05 EDT
Subject: Re: Stereometry by computer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In a message dated 6/8/00 12:38:48 PM, bozzola-at-siu.edu writes:

} A user of our facility is interested in making height measurements
} from specimens viewed in the SEM. I am aware of the conventional way
} of doing this: stereo pairs and optical viewer (with stereometer
} parallax corrections). Is there a more modern (computerized) way of
} doing this, say with anaglyphs?


There was a thread recently on measurement of stereo pairs by
computer-matching of points. Several software packages (one of the is Fovea
Pro - http://members.aol.com/FoveaPro) were mentioned as having this
capability. Whether or not the two images are put together as an anaglyph is
unimportant.





From: David Bentley :      dlb-at-u.arizona.edu
Date: Thu, 08 Jun 2000 11:39:28 -0700
Subject: Re: Help? Cleaning Optical Lenses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Just a couple of additional comments to Barbara's procedure.

A Zeiss axiom: If the lens is not dirty, then cleaning it will never
improve it, it only risks damaging it. Do the least amount possible to
return the surface to clean. If the surface is only dusty, & puffing it
removes the problem, stop there. I look at the lens being cleaned, using
the ocular turned upside down as a loupe, at each step of the process.
When it's clean, you're done. Usually only external surfaces are
contaminated and need frequent or extensive cleaning. Internal lenses, a
dusting often suffices.

In dealing with a dissecting microscope with a zoom magnification
capability as opposed to one with click stops or a fixed magnification, be
very careful about disturbing the positional settings of the internal
lenses. Any changes to these settings will alter magnification for the
each of the two eyes and make an image that can't be justified in the
brain. This will require a service engineer to remedy. For heavily filmed
lenses, use short or broken cotton tipped applicator and get in the best
you can. Varying the zoom control will move the lenses up and down and
maybe give you enough room to work.

As Barbara wrote, removing loose dust is essential to preserving lens
quality. Lenses are coated with various coatings and these are easily
damaged. Puffers from camera stores are good. The red bulb ear syringe
for babies is excellent and usually readily available. I do use dusters
and compressed air although both are frowned upon, by some, as possibly
damaging lens coatings. If you use a duster, don't shake or tip it while
dusting the lens. This will expel liquid from the can and contaminate the
lens worse.

A good lens cleaner that is easy to get is Sparkle Glass Cleaner available
from Ace hardware, and grocery stores (No financial interests). Never use
Windex. It contains oils that will coat the lens.

As much as possible use the cotton tipped applicators rather than lens
tissues. Unless you wear latex or polyethylene gloves, finger oils will
get onto the tissue and be transferred to the lens. Another reason to shy
away from the tissue is the tendency to scrub the lens. I agree with
Barbara, NEVER, NEVER Kimwipes, I've been told from many sources they
contain many silica strands, and will scratch the delicate optical
coatings. On expensive lenses, make a single pass, using minimal pressure
straight across the lens, with the applicator, rolling the stick between
your fingers to present a fresh cotton surface to the lens and picking up
and getting the dust away from the lens. Alternate between wet (either
with lens cleaner or condensed breath (open mouth, no spit please) and dry
cotton. (Yes, it is often necessary to use a swirling motion on oculars
because of the amount of oil contamination from the eye lashes.)

It is best not to try to clean filters, there are a few recommended
procedures but all potentially damage the very precise and thin coatings of
the filter compromising its performance. (If anyone disagrees or has a
favorite procedure for filter cleaning, contact me off list, I'd like to
hear about it.)

Use of solvents or disassembling objectives or multiple lens stacks is
best left to the service engineer. Lenses use a variety of air and cement
interfaces to achieve resolution and aberration correction. Altering these
by dissolving the cement will degrade the lens. Getting the small lenses
back in exactly the same position is also very, very difficult and not for
the faint hearted.

There is a microscope repair workshop in the initial planning stages for
the Long Beach MSA meeting next year. It will target K-12 school teachers,
but will hopefully address many levels. If members of the list will be at
the meeting and are interested in attending this workshop to: 1. learn
basic repairs for their microscopes. 2. get ideas for their outreach
program. or 3. assist with putting on the workshop. Please e-mail me
telling me of your interest or asking for more information.

If you have any questions feel free to contact me off line.




From: Soumitra Ghoshroy :      ghoshroy-at-nmsu.edu
Date: Thu, 8 Jun 2000 12:11:37 -0700
Subject: RE: plant vesicles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jim is absolutely right about LR white. But instead of increasing the GA I
will probably increase paraformaldehyde conc. to 4%.

Good luck,

Soumitra


} } Hi,
} }
} } I'm trying to do immunogold studies on a membrane protein in a preparation
} } of plant vesicles (from minus 80oC storage). I did an Epon embedding with
} } standard procedure (glut and osmium) to check the ultrastructure and got
} } decent-looking vesicles, but side by side an LR White prep (0.5% glut, 2%
} } paraformaldehyde, without osmium) looks like a disaster, with granular
} } material, vague membranish-like structures and whorls of membranes. Does
} } anyone have any tips for LR White preps of vesicles, or any other ideas
} } for embedding (resins, fixing, dehydration, embedding) to favour the
} } immunogold process.
} }
} } Thanks,
} }
} } Mark
} }
} }
} }
} }
} }
} }
} } ********************************************
} } Mark West,
} } Unidad de Microscopia Electronica,
} } (Electron Microscopy Unit)
} } Instituto de Fisiologia Celular,
} } Universidad Nacional Autonoma de Mexico,
} } 04510 Mexico D.F.
} }
} } tel (unidad/lab) *(525) 622 5610*
} } (casa/home) (525) 619 3020
} } Fax (525) 616 2282
} } ********************************************
} }




*****************************************************************
Soumitra Ghoshroy Ph.D.
Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3600
Fax: 505-646-5665
e-mail: ghoshroy-at-nmsu.edu
http://confocal.nmsu.edu/eml






From: Tamara Howard :      howard-at-cshl.org
Date: Thu, 8 Jun 2000 15:21:15 -0400 (EDT)
Subject: Re: Help? Cleaning Optical Lenses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Thu, 8 Jun 2000, David Bentley wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
text deleted
}
} It is best not to try to clean filters, there are a few recommended
} procedures but all potentially damage the very precise and thin coatings of
} the filter compromising its performance. (If anyone disagrees or has a
} favorite procedure for filter cleaning, contact me off list, I'd like to
} hear about it.)
}
Please respond on-line! I'm sure that I'm not the only other person who
would be interested in this!

Tamara Howard
CSHL






From: Ann-Fook Yang (Ann-Fook Yang) :      yanga-at-Em.agr.ca
Date: Thu, 08 Jun 2000 16:44:55 -0400
Subject: Re: plant vesicles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Try treat the glut- & Os-fixed sections you have with sodium metaperiodate before doing immunolabeling. It may work.

Reference:
M. Bendayan 1989. Protein A-gold and Protein G-gold postembedding immunoelectron microscopy. In Colloidal Gold: Principles, Methods and Applications Vol.1 Academic Press.



Ann Fook Yang
EM Unit,
Eastern Cereal and Oilseed Research Centre,
Rm 2091, K.W. Neatby Bldg.,
Central Experimental Farm,
Ottawa, Ontario, Canada K1A 0C6

Phone: 613-759-1638
Fax; 613-759-1701

} } } Mark West {mwest-at-ifcsun1.ifisiol.unam.mx} 06/07 11:50 AM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hi,

I'm trying to do immunogold studies on a membrane protein in a preparation
of plant vesicles (from minus 80ĽC storage). I did an Epon embedding with
standard procedure (glut and osmium) to check the ultrastructure and got
decent-looking vesicles, but side by side an LR White prep (0.5% glut, 2%
paraformaldehyde, without osmium) looks like a disaster, with granular
material, vague membranish-like structures and whorls of membranes. Does
anyone have any tips for LR White preps of vesicles, or any other ideas
for embedding (resins, fixing, dehydration, embedding) to favour the
immunogold process.

Thanks,

Mark






********************************************
Mark West,
Unidad de Microscopia Electronica,
(Electron Microscopy Unit)
Instituto de Fisiologia Celular,
Universidad Nacional Autonoma de Mexico,
04510 Mexico D.F.

tel (unidad/lab) *(525) 622 5610*
(casa/home) (525) 619 3020
Fax (525) 616 2282
********************************************







From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Fri, 9 Jun 2000 08:48:46 GMT+1200
Subject: Re: JEOL 840 SEM Vacuum Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sorry, I don't have any useful suggestions, and for that reason, I'd
rather see responses posted to the list, if possible, as this gives
me a wonderful and otherwise unavailable opportunity to learn from
the experience of others.

cheers

rtch







} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} We have JEOL 840 SEM that has develop a leak at the specimen chamber door
} interface and will not hold a vacuum. We have changed the o-ring around the
} door, look for nicks, scratches any other damage around the door interface, none
} found. We have also check the roughing valve (V4, LV3 and the pressure relief
} valve) for leaks. Pressure holds in the gun area. The large door is only opened
} when we have to replace or add a new attachment to the chamber and thats about
} once a year. We have a vacuum specimen exchange port to enter and retrieve
} samples. We have also check that seal and it also checks out OK.
} We have had 2 JEOL engineers take a look at the system and at this point no luck
} finding the cause or solution. You can respond to me off-line at
} bruce.arey-at-pnl.gov
} Thanks
}
} Bruce W. Arey
}
} PNNL-Battelle
} Associate Scientist
} Microstructural Characterization
} bruce.arey-at-pnl.gov
} 509-376-3363
} fax 509-376-6308
}
}
}
}
Ritchie Sims Phone : 64 9 3737599 ext 7713
Department of Geology Fax : 64 9 3737435
The University of Auckland email : r.sims-at-auckland.ac.nz
Private Bag 92019
Auckland
New Zealand




From: hpadams :      hpadams-at-bcm.tmc.edu
Date: Thu, 8 Jun 2000 15:59:48 -0500
Subject: RE: plant vesicles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mark, you didn't say how you post-stained after you immunolabeled. This step
can make all the difference in the world. For LRW, I prefer 3-4 secs in
saturated UA in 50% etoh, wash, followed by about 15secs in lead citrate. I
find a major difference (inferior) when I use aqueous UA. Also, going too long
in any of the solutions will give poor results.

} ===== Original Message From Mark West {mwest-at-ifcsun1.ifisiol.unam.mx} =====
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hank Adams
Manager
Integrated Microscopy Core
Molecular and Cellular Biology
Baylor College of Medicine
Houston, Tx 77030





From: David Joswiak :      joswiak-at-orca.astro.washington.edu
Date: Thu, 8 Jun 2000 15:43:51 -0700 (PDT)
Subject: Re: Stereometry by computer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John - I routinely make height measurements of 5 - 10 um interplanetary
dust particles using the objective focus knob on our JEOL 1200EX in STEM
mode. The procedure is simply to focus on the top of the particle and
note the number of steps (or count the number of clicks of the knob)
required when refocussing on the substrate next to the particle. By
calibrating the objective focus step size with a known standard, you can
routinely measure particle heights. In principle, a similar procedure
should be possible on an SEM, depending on how the objective focus is
configured. I don't think you can expect high accuracy with this method,
however, so it may not be applicable to your needs.

Dave Joswiak
Dept. of Astronomy
University of Washington
Seattle, WA 98195
joswiak-at-astro.washington.edu

On Thu, 8 Jun 2000, John J. Bozzola wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} A user of our facility is interested in making height measurements
} from specimens viewed in the SEM. I am aware of the conventional way
} of doing this: stereo pairs and optical viewer (with stereometer
} parallax corrections). Is there a more modern (computerized) way of
} doing this, say with anaglyphs?
}
} Thank you.
}
} John B.
}
}
} ####################################################################
} John J. Bozzola, Ph.D., Director
} Micro-Imaging and Analysis Center
} 750 Communications Drive - MC 4402
} Southern Illinois University
} Carbondale, IL 62901 U.S.A.
} Phone: 618-453-3730
} Fax: 618-453-2665
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/departments/shops/cem.html
} ####################################################################
}
}





From: Robert Mixon :      mixonr-at-ohsu.edu
Date: Thu, 08 Jun 2000 15:59:22 -0700
Subject: RE: Cleaning optical lenses and filters.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tamara et al : One superior method for cleaning sensitive optical components such as filters and front surfaced telescope mirrors is as follows: 1) gently puff of extraneous dust. 2) gently float the surface with deionized water. 3) if necessary to remove dirt film use a q-tip soaked in a non-ionic surfactant like a 0.01% solution of triton-x 100 (IMPORTANT DO NOT RUB ON THE SURFACE WITH FORCE BUT SIMPLY LET THE WEIGHT OF THE WET Q-TIP(S) BE THE ONLY "DOWNWARD PRESSURE" ON THE SURFACE OF THE GLASS. Such gently techniques such as these should help mitigate damage.





From: Robert Mixon :      mixonr-at-ohsu.edu
Date: Thu, 08 Jun 2000 16:03:20 -0700
Subject: RE: Cleaning optical lenses and filters.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tamara et al : (excuse mispelling in first copy sent) One superior method for cleaning sensitive optical components such as filters and front surfaced telescope mirrors is as follows: 1) gently puff off extraneous dust. 2) gently float the surface with deionized water. 3) if necessary to remove dirt film use a q-tip soaked in a non-ionic surfactant like a 0.01% solution of triton-x 100 (IMPORTANT DO NOT RUB ON THE SURFACE WITH FORCE BUT SIMPLY LET THE WEIGHT OF THE WET Q-TIP(S) BE THE ONLY "DOWNWARD PRESSURE" ON THE SURFACE OF THE GLASS. Such gently techniques such as these should help mitigate damage.





From: Caroline Schooley :      schooley-at-mcn.org
Date: Thu, 8 Jun 2000 16:02:44 -0700
Subject: Re: LRWhite flat embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} I remember a message some time ago, I think, suggesting a way to flat
} embed in silicon molds using LRWhite resin. How the top was sealed
} escapes me. Could anyone with a better filing system and/or memory
} please send me a message or a reference on this.

Peter Bond {P.Bond-at-plymouth.ac.uk}

} Pete -

If you have a silicon mold that isn't permeable to LR White, I'd like to
know who makes it! There is a teflon flat mold available from Ted Pella
that works well with LR White; it's their own product. It's sealed with
Thermanox coverslips, so you might try them with your mold.


Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: Dopeyee-at-aol.com
Date: Thu, 8 Jun 2000 20:28:48 EDT
Subject: LEAK IN JEOL

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


GET A HOLD OF A LEAK DETECTOR AND FIND THE TRUE AREA THAT IS LEAKING

I DO BELEIVE THE MANUFACTURER SHOULD BE OF MORE HELP

THEY HAVE TO FIX IT!!!!




From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Thu, 08 Jun 2000 22:11:20 -0500
Subject: LR White flat embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Peter Bond wrote:
==============================================================
I remember a message some time ago, I think, suggesting a way to flat embed
in silicon molds using LRWhite resin. How the top was sealed escapes me.
Could anyone with a better filing system and/or memory please send me a
message or a reference on this.
==============================================================
This might have been an old posting of mine. Since UV transparency is
usually required, we are talking about clear UV transparent silicone
embedding molds and we recommend a slight "overfilling" of the cavities.
Then when all cavities are filled, take another mold, just like the first
one, and place it on top, bottom side down, cavity side up. The capillary
action will result in a sealing out of any oxygen.

When the UV curing is complete, the top mold can be easily separated and
since the cavity side is still unused, no wear and tear has been put on it
in terms of taking away any of its lifetime.

Information about these transparent-to-UV embedding molds, and their use,
can be found on the SPI website given below.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
============================================






From: Oldrich Benada :      benada-at-biomed.cas.cz
Date: Fri, 9 Jun 2000 09:19:55 +0200
Subject: A hole in the EDS spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear colleagues,
We have a problem with spectra recording in our EDAX DX4 system
mounted on Philips CM12/STEM. There is a hole in the spectrum between
0.5keV and 3.5keV.
With service engineer, we have borrowed all the boards in electronics
and exchange our boards with the borrowed ones. The hole in the
spectrum remains. We have also dismounted detector from the EM column
and the Be window was checked for contamination. The window was clean
and intact.
Please, can anybody give us any hints how to solve our problem? Many
thanks in advance for any comment.
Oldrich Benada

Our system specification:
Philips CM12/STEM
EDAX DX4 system with 184 preamp. based on win3.11





+-----------------------------------+
Oldrich Benada
Acad. Sci. CR
Institute of Microbiology
Laboratory of electron microscopy
Videnska 1083
CZ - 142 20 Prague 4 - Krc
Czech Republic
+------------------------------------+
Phone: +420-2-4752399
Fax: +420-2-4715743
WEB:http://www.biomed.cas.cz/mbu/lem113/lem.htm




From: Ulrike Zeile :      ulrike.zeile-at-bal-tec.com
Date: Fri, 9 Jun 2000 09:44:26 +0200
Subject: AW: service

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Mr.Patel,

Balzers Union renamed to BAL-TEC AG in 1992. But the product
line was maintained and new developments are carried out.
For additional information have a look at our website www.bal-tec.com.
There are several coating systems and freeze fracture systems of this older
types in use.

Our representative in USA / Canada:

Techotrade International
7 Perimeter Road
Manchester, NH 03103-3343
T: +1 603 622-5011

Our representative will contact you directly to assist with
further information.

-----Ursprźngliche Nachricht-----
Von: Rajesh Patel [mailto:rpatel-at-UMDNJ.EDU]
Gesendet: Mittwoch, 31. Mai 2000 19:42
An: microscopy-at-sparc5.microscopy.com
Betreff: service


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America



We need service on our Balzer 500 freeze fractrure unit,
and was wondering who services them. Anyone we can contact?

Rajesh Patel
Robert Wood Johnson Medical School
Dept. of Pathology
Electron Microscopy Lab
675 Hoes Lane
Piscataway, NJ 08854

(732)235-4648
rpatel-at-umdnj.edu




From: Dr Malcolm Roberts :      malc-at-rock.ru.ac.za
Date: Fri, 09 Jun 2000 11:53:08 +0200
Subject: Carbonate standards - bizarre behaviour

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


G'day Folks,
I acquired some carbonate standards a while ago from the Smithsonian
Insitute. Namely, Calcite, Siderite, Dolomite and Strontianite.
Interestingly, when admiring them with our ED system, there appeared an
anomalous peak around about where B is supposed to be. This is not a
detector artifact as it is specific only to these materials. I just
wondered if anyone else had noticed the same thing, or whether anyone
has any clues on why this should be occuring. The programme suggests
that there is about 50 wt % B in these things which seems unbelievable
considering the compositions supplied by Washington. Ideas?
Cheers,
Malc.

--
Dr MP Roberts Phone: [+27](0)46 603 8313
Dept of Geology Fax: [+27](0)46 622 9715
Rhodes University Cell: 083 4060 262 (usually off)
6140 Grahamstown e-mail: malc-at-rock.ru.ac.za
SOUTH AFRICA






From: HARRISm-at-esm-semi.co.uk
Date: Fri, 09 Jun 2000 10:12 +0000 (GMT)
Subject: RE: JEOL840 SEM VACUUM PROBLEMS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



It may not help but .....

With regard to the vacuum problems encountered by Bruce , without
being familiar with the instrument we encountered a ' similar '
problem and were misled into thinking a chamber door leak existed by
an over zealous leak detection unit .
The problem on our sem actually turned out to be a hairline crack in a
metal bellows attached to a valve that appeared OK . The vaccum would
not exceed a certain level in pumpdown in its final stages .

I believe these bellows were often a source of problems on older
instruments .

M.HARRIS harrism-at-esm-semi.co.uk
ESM LTD
South Wales , U.K .





From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Fri, 9 Jun 2000 08:51:11 -0400
Subject: RMC microtomes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,
I'm posting this for a friend who is not on-line.

Does anyone out there have a used RMC MT-7000 for sale? My friend just
moved to a lab that had an MT-5000 which she was told was operational. A
major exaggeration. She is an RMC loyalist, and desparately wants another,
but has a limited budget.

Please contact: Linda Burg Friedman at Columbia P & S at (212)305-9047

Thanks,
Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Confocal Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Steve Chapman :      protrain-at-emcourses.com
Date: Fri, 9 Jun 2000 14:34:56 +0100
Subject: Re: JEOL 840 SEM Vacuum Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi

The specimen chamber is a vast area with many potential leak sources.

The door is the most likely if it is being opened on a regular basis,
however do not forget that the stage drives pass through the door and they
are being used all of the time.

If you are sure the door "O" ring is OK check the stage drive feed throughs
as they may be the source of your problem?

Steve Chapman
Senior Consultant Protrain
For consultancy and professional training in EM world wide
Tel 44+ 1280 814774 Fax 814007
www.emcourses.com





From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Fri, 9 Jun 2000 09:05:24 -0400
Subject: Re: TEM: Set-up, operating and facilities costs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rick,
YOu don't mention make or modle of TEM, but i can tell you that my service
contract (2 preventive maintenance call, unlimited service calls, parts,
labor) costs around $15,500/year.
You will need a water supply for cooling and electrical work to bring in a
dedicated line. You will need a dakrroom with running water and a temp.
control valve to process the film. If you wish to make photographic prints
of your negatives, you will also need a point source enlarger (Durst was
the best, but they are hared to find, Omega used to make one, too) and
either pans (slow and painful) or a rapid processor.
Budgeting for supplies is a tough call...it depends on your usage.
If you find someone with an EM background to run it, its should't take them
long to learn the individual instrument. Training someone from ground zero
could take months to get dthe person on his/her way.
Ancillary euqipment: it depends on what you will be doing, biological or
materials, embedding/sectioning or particulates, negataive staining or
metal shadowing.

A really good source to check is Audrey Glauert's Practical Methods in
Electron Microscopy Vol. 4: Design of the Electron Microscope Laboratory,
by Ronald H. Alderson, North-Holland publishers, 1975.

If you wish to contact me off-line, I can go over my lab's operating budget
with you. I run a basic EM Core Facility here at (what used to be called)
Cornell Medical College.

Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Confocal Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Arey, Bruce W :      bruce.arey-at-pnl.gov
Date: Fri, 09 Jun 2000 07:05:10 -0700
Subject: RE: LEAK IN JEOL

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


JEOL has been very helpful and they are somewhat limited on what they can do to
this instrument. The 840 has radioactive particles inside the chamber so we have
to be careful each time we open the system up plus there hands on is limited to
just oversight. I am in the process of trying to find a leak detector on site
that can be used on radioactive system. But we also cannot pump down the chamber
enough to use a leak detector. So we are in the process of making a plate that
will fit over the door. We have taken off all our accessories off the chamber
(EDS, OIM, BSE) and have pressurized the system and have found some evidence of
a leak on the left hand side on the door, we have changed the o-ring and still
the seal leaks on this side of the door. We have polished the door seal to make
sure there is no major scratches or marks. We are pretty confident that specimen
exchange port is OK we can pump this down and we see no signs of a leak. Why are
we focusing our attention on the large o-ring and chamber door. We can clean
the o-ring with alcohol (methanol or ethanol) and we can rough pump the chamber
out and put the chamber on the high vac system but after a period of time the
alcohol dries out and the system shuts down too a poor vacuum (overnight). Then
we try to rough pump the chamber and we cannot go more than 20-30 mamps
difference on the pirani gauge. We have done this several times with the same
results. We are going to try to pressurize the system with He and try He sniffer
to help us maybe pin point the leak. Thanks to all who have responded with
suggestion on finding the leak and if we are successful in finding the leak I
will post the finding on the server. Again JEOL has been very responsive in
trying to find the leak and they are somewhat limited on this instrument do to
its environment. Any other suggestion are welcomed.


Bruce W. Arey

PNNL-Battelle
Associate Scientist
Microstructural Characterization
bruce.arey-at-pnl.gov
509-376-3363
fax 509-376-6308


----------
From: "Dopeyee-at-aol.com"-at-sparc5.microscopy.com
Sent: Thursday, June 8, 2000 4:28 PM
To: Microscopy-at-sparc5.microscopy.com
Subject: LEAK IN JEOL

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America
To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
-----------------------------------------------------------------------.


GET A HOLD OF A LEAK DETECTOR AND FIND THE TRUE AREA THAT IS LEAKING

I DO BELEIVE THE MANUFACTURER SHOULD BE OF MORE HELP

THEY HAVE TO FIX IT!!!!





From: Huggins, Bradley J :      HUGGINBJ-at-bp.com
Date: Fri, 9 Jun 2000 09:05:55 -0500
Subject: RE: JEOL 840 SEM Vacuum Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I recall many years ago we had a similar vacuum leak on our 840A that was
related to the specimen exchange port mechanism. The moving parts in the
sliding door mechanism had become slightly miss-aligned. We ended up
disassembling that mechanism, installing new o-rings and lubricating with
Apiezon L. Problem gone.
Hope your leak is that easy!
Good Luck
Brad Huggins
BP Amoco, Naperville

} ----------
} From: Arey, Bruce W[SMTP:bruce.arey-at-pnl.gov]
} Sent: Thursday, June 08, 2000 9:24 AM
} To: 'Microscopy-at-msa.microscopy.com'
} Subject: JEOL 840 SEM Vacuum Problem
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} We have JEOL 840 SEM that has develop a leak at the specimen chamber door
} interface and will not hold a vacuum. We have changed the o-ring around
} the
} door, look for nicks, scratches any other damage around the door
} interface, none
} found. We have also check the roughing valve (V4, LV3 and the pressure
} relief
} valve) for leaks. Pressure holds in the gun area. The large door is only
} opened
} when we have to replace or add a new attachment to the chamber and thats
} about
} once a year. We have a vacuum specimen exchange port to enter and retrieve
} samples. We have also check that seal and it also checks out OK.
} We have had 2 JEOL engineers take a look at the system and at this point
} no luck
} finding the cause or solution. You can respond to me off-line at
} bruce.arey-at-pnl.gov
} Thanks
}
} Bruce W. Arey
}
} PNNL-Battelle
} Associate Scientist
} Microstructural Characterization
} bruce.arey-at-pnl.gov
} 509-376-3363
} fax 509-376-6308
}
}
}




From: milesd-at-us.ibm.com
Date: Fri, 9 Jun 2000 10:24:19 -0400
Subject: RE: Cleaning optical lenses and filters.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have always understood that deionized water would etch metal,
that being the reason for PVC pipe being used to deliver it. If this
is true, wouldn't that damage the first surface mirrors?

Darrell





From: ComCryLab1-at-aol.com
Date: Fri, 9 Jun 2000 10:31:59 EDT
Subject: LM: alignment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


David,
We have a B&L StereoZoom (0.7X - 3X) microscope which we
intend to donate to the local school system
(Computer VoTech Dept) for use in examining the circuit patterns
on silicon wafers that I have donated to them.

My problem is that the microscope is not functioning properly.
The two optical paths are not synchronized during zooming and
therefore the image tends to rotate or it falls out of focus. We have
tried to repair it in-house but we were unsuccessful. Knowing that
that we would create additional problems, we did not disturb the
lens or prism sub-assembly.

Is there a set of maintenance instructions or a tech manual available
to help us align this scope?

Thanks
Mike Urbanik
Commercial Crystal Labs
Naples, FL
www.crystalguru.com




From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Fri, 9 Jun 2000 08:46:11 -0600
Subject: Re: Stereometry by computer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John,

you may want to try the method below if your samples show very large
structures. One of the nice features of an SEM is it's large depth of
focus, unfortunately in many cases this prevents you from using the
technique mentioned below. You can use a stereo technique to calculate a
surface profile (see for example the stereo module on our web site or
other stereo applications). Typical results from stereo images have a
height resolution of about 1/10 the of the lateral resolution (in other
words: if your lateral resolution is 1 micron between pixels, the height
resolution will be on the order of 10 microns). This can be improved by
sub-pixel interpolation but gives you an order of magnitude for the
resolution.

Michael

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: David Joswiak [mailto:joswiak-at-orca.astro.washington.edu]
Sent: Thursday, June 08, 2000 4:44 PM
To: John J. Bozzola
Cc: Microscopy-at-sparc5.microscopy.com


John - I routinely make height measurements of 5 - 10 um interplanetary
dust particles using the objective focus knob on our JEOL 1200EX in STEM
mode. The procedure is simply to focus on the top of the particle and
note the number of steps (or count the number of clicks of the knob)
required when refocussing on the substrate next to the particle. By
calibrating the objective focus step size with a known standard, you can
routinely measure particle heights. In principle, a similar procedure
should be possible on an SEM, depending on how the objective focus is
configured. I don't think you can expect high accuracy with this
method,
however, so it may not be applicable to your needs.

Dave Joswiak
Dept. of Astronomy
University of Washington
Seattle, WA 98195
joswiak-at-astro.washington.edu

On Thu, 8 Jun 2000, John J. Bozzola wrote:

}
------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
-----------------------------------------------------------------------.
}
}
} A user of our facility is interested in making height measurements
} from specimens viewed in the SEM. I am aware of the conventional way
} of doing this: stereo pairs and optical viewer (with stereometer
} parallax corrections). Is there a more modern (computerized) way of
} doing this, say with anaglyphs?
}
} Thank you.
}
} John B.
}
}
} ####################################################################
} John J. Bozzola, Ph.D., Director
} Micro-Imaging and Analysis Center
} 750 Communications Drive - MC 4402
} Southern Illinois University
} Carbondale, IL 62901 U.S.A.
} Phone: 618-453-3730
} Fax: 618-453-2665
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/departments/shops/cem.html
} ####################################################################
}
}





From: Tracey M. Pepper :      tpepper-at-iastate.edu
Date: Fri, 09 Jun 2000 12:29:09 -0500
Subject: LR White flat embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This may sound like "bucket science" but we've used the disposable aluminum
weighing dishes to flat embed material. If you fill the bottom tray about
1/2 full and set another tray inside it, press gently until a little LR
white oozes up the sides, what remains in the middle will be protected from
the air and should polymerize nicely. (Maybe we've just been lucky!)
good luck!

Tracey Pepper
Supervisor
Bessey Microscopy Facility
Iowa State University
ph: 515-294-3872
fax: 515.294.1337





From: Jim McGee :      mcgee-at-geol.sc.edu
Date: Fri, 09 Jun 2000 13:53:37 -0400
Subject: Re: Carbonate standards - bizarre behaviour

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Malcolm:

Some Chlorine-L lines, as well as Sr lines have nearly identical
energy/wavelength as B Ka. I don't remember the chemistry of these
standards offhand, but I know some of them were Sr-bearing and some may also
have Cl. These are likely the peaks you are seeing at the low end.

Jim
--
^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
James J. McGee (email: jmcgee-at-sc.edu)
Department of Geological Sciences
University of South Carolina
Columbia, SC 29208

Tel: 803-777-6300 Fax: 803-777-6610


Dr Malcolm Roberts wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} G'day Folks,
} I acquired some carbonate standards a while ago from the Smithsonian
} Insitute. Namely, Calcite, Siderite, Dolomite and Strontianite.
} Interestingly, when admiring them with our ED system, there appeared an
} anomalous peak around about where B is supposed to be. This is not a
} detector artifact as it is specific only to these materials. I just
} wondered if anyone else had noticed the same thing, or whether anyone
} has any clues on why this should be occuring. The programme suggests
} that there is about 50 wt % B in these things which seems unbelievable
} considering the compositions supplied by Washington. Ideas?
} Cheers,
} Malc.
}
} --
} Dr MP Roberts Phone: [+27](0)46 603 8313
} Dept of Geology Fax: [+27](0)46 622 9715
} Rhodes University Cell: 083 4060 262 (usually off)
} 6140 Grahamstown e-mail: malc-at-rock.ru.ac.za
} SOUTH AFRICA









From: bobrob-at-uswest.net
Date: Fri, 09 Jun 2000 11:49:20 -0700
Subject: SEM: JEOL 840 Vacuum Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


If the instrument has been physically moved recently, this can sometimes

hasten a crack in these stainless steel "flex" vacuum lines. I have
found
these to develop mostly at the end of the tube where it has been flared
for
the fitting connector. Check the roughing lines first.

Happy hunting...

Bob Roberts
EM Lab Services, Inc
2409 S. rural Rd Suite C
Tempe, Arizona 85282
480.967.3946





From: Huggins, Bradley J :      HUGGINBJ-at-bp.com
Date: Fri, 9 Jun 2000 12:18:33 -0500
Subject: RE: A hole in the EDS spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Oldrich
Need a little more information.
Do you actually have spectral information in the spectrum at energies { 0.5,
or is there possibly, only noise in this low energy range? If the signal
present in your spectra at these very lowest energies is just noise-like
signal. Then it is possible that you have a severe alignment problem with
the EDS detector/specimen geometry. A combination of high noise (due to
vibration or other sources) and poor line of sight with the detector window
(resulting in detection of only the higher energy x-rays ) could give you
this "hole in the EDS spectrum" effect. A miss-aligned detector, and/or a
detector making contact with the internal parts of the microscope might
create this situation. Does the information in the low energy region
correlate with the specimen composition? Does the high energy signal
correlate with the specimen composition?

} ----------
} From: Oldrich Benada[SMTP:benada-at-biomed.cas.cz]
} Sent: Friday, June 09, 2000 2:19 AM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: A hole in the EDS spectrum
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear colleagues,
} We have a problem with spectra recording in our EDAX DX4 system
} mounted on Philips CM12/STEM. There is a hole in the spectrum between
} 0.5keV and 3.5keV.
} With service engineer, we have borrowed all the boards in electronics
} and exchange our boards with the borrowed ones. The hole in the
} spectrum remains. We have also dismounted detector from the EM column
} and the Be window was checked for contamination. The window was clean
} and intact.
} Please, can anybody give us any hints how to solve our problem? Many
} thanks in advance for any comment.
}
} Oldrich Benada
}
} Our system specification:
} Philips CM12/STEM
} EDAX DX4 system with 184 preamp. based on win3.11
}
}
}
}
}
} +-----------------------------------+
} Oldrich Benada
} Acad. Sci. CR
} Institute of Microbiology
} Laboratory of electron microscopy
} Videnska 1083
} CZ - 142 20 Prague 4 - Krc
} Czech Republic
} +------------------------------------+
} Phone: +420-2-4752399
} Fax: +420-2-4715743
} WEB:http://www.biomed.cas.cz/mbu/lem113/lem.htm
}




From: hard-at-acsu.buffalo.edu
Date: Fri, 09 Jun 2000 16:18:05 -0500
Subject: Course Announcement I

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Course Announcement

Title: Optical Microscopy and Imaging in the Biomedical Sciences

When: October 11 - October 19, 2000

Where: Marine Biology Laboratory, Woods Hole, MA, USA

Tuition: $2150 (Includes room and board, text, handouts, supplies)

Application Deadline: August 1, 2000

Admission application and information:
Carol Hamel, Admissions Coordinator
Marine Biological Laboratory
7 MBL Street
Woods Hole, MA 02543-1015
(508) 289-7401
Internet: admissions-at-mbl.edu
WWW: http://www.mbl.edu (Application forms available via Adobe Acrobat)

Course Director: Colin S. Izzard, State University of New York -at- Albany
Phone: [518] 442 - 4367
EMail: csizzard-at-csc.albany.edu

Course Description:

For Whom:
Designed primarily for research scientists, physicians, postdoctoral
trainees and advanced graduate students in animal, plant, medical and
material sciences. Non-biologists seeking a comprehensive introduction to
microscopy and video-imaging will benefit greatly from this course as well.
There are no specific prerequisites, but an understanding of the basic
principles of optics is desirable. Limited to 24 students.

The eight day course consists of lectures, laboratory demonstrations,
exercises and discussions that will enable the participant to obtain and
interpret microscope images of high quality, to perform quantitative optical

measurements, and to produce photographic and video records for
documentation
and analysis.

Topics to be covered include:
principles of microscope design and image formation
bright field, dark field, phase contrast, polarized light,
differential
interference contrast, interference reflection, and fluorescence
microscopy
confocal scanning microscopy, multiphoton excitation fluorescence
microscopy and image deconvolution
digital image restoration and 3-D reconstruction
video imaging, recording, enhancement, and intensification
analog and digital image processing and analysis
fluorescent probes and ratiometric-imaging
laser tweezers and laser scissors

Applications to live cells will be emphasized; other specimens will be
covered as well.

Students will have direct hands-on experience with state-of-the-art
microscopes, video cameras, recorders and image processing equipment
provided
by major optical and electronics companies. Instruction will be provided by
experienced staff from universities and industry.

Students are encouraged to bring their own biological (primary
cultures, cell lines, etc.) and material specimens and to discuss individual

research problems with the faculty.





From: David Cockayne :      david.cockayne-at-materials.oxford.ac.uk
Date: Fri, 9 Jun 2000 17:38:24 -0500
Subject: University of Oxford: Job Vacancies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


University of Oxford: Job Vacancies



DEPARTMENT OF MATERIALS



Postdoctoral Research Assistant - Electron Microscopy of Crystalline
Materials



Salary Ł16,286 - Ł24,479 p.a.

A three-year position is available in a research group being developed by
Professor David Cockayne FRS for the study of a range of materials using
electron diffraction, electron microscopy (EM) and modelling techniques.
Quantitative microscopy and materials modelling will refine structural
models of technologically important materials. Extensive expertise in EM,
diffraction, the preparation of crystalline materials for microscopy, and
strong computing skills, are essential. Expertise in microscopy of
semiconductors including quantum dots and computational techniques for
image simulation would be an advantage. The Department has outstanding EM
and modelling facilities. Please quote ref. DJ00/12.

Applications including a curriculum vitae, list of publications and the
names and addresses of three referees should be sent to The Administrator,
Department of Materials, University of Oxford, Parks Road, Oxford, OX1
3PH, from whom further particulars are available. The closing date for
applications is 14 June 2000.











From: jim :      jim-at-proscitech.com.au
Date: Sat, 10 Jun 2000 13:56:13 +1000
Subject: RE: LEAK IN JEOL

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have found the use of clear silcone sealant (as used for windows etc) quite
effective to temporally fix or help to isolate large and medium vacuum leaks.
In many instance it can be applied externally over various fittings. Screw
holes are best covered first with a little tape, to facilitate the later the
removal of the dry silicone. Silicone outgases a fair bit for some hours, so
only major leaks can be determined immediately after applying the silicone.
The method seems crude but is effective to eliminate numerous fittings as the
source of a leak. I once operated a TEM for several months with a split
stainless bellow, patched with a smear of silicone sealant.
I don't suggest the use of that sealant on a permanent basis or in ion gutter
pumped parts of a column.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Saturday, June 10, 2000 12:05 AM, Arey, Bruce W [SMTP:bruce.arey-at-pnl.gov]
wrote:
}
}
} JEOL has been very helpful and they are somewhat limited on what they can do
} to
} this instrument. The 840 has radioactive particles inside the chamber so we
} have
} to be careful each time we open the system up plus there hands on is limited
} to
} just oversight. I am in the process of trying to find a leak detector on site
} that can be used on radioactive system. But we also cannot pump down the
} chamber
} enough to use a leak detector. So we are in the process of making a plate
} that
} will fit over the door. We have taken off all our accessories off the chamber
} (EDS, OIM, BSE) and have pressurized the system and have found some evidence
} of
} a leak on the left hand side on the door, we have changed the o-ring and
} still
} the seal leaks on this side of the door. We have polished the door seal to
} make
} sure there is no major scratches or marks. We are pretty confident that
} specimen
} exchange port is OK we can pump this down and we see no signs of a leak. Why
} are
} we focusing our attention on the large o-ring and chamber door. We can clean
} the o-ring with alcohol (methanol or ethanol) and we can rough pump the
} chamber
} out and put the chamber on the high vac system but after a period of time the
} alcohol dries out and the system shuts down too a poor vacuum (overnight).
} Then
} we try to rough pump the chamber and we cannot go more than 20-30 mamps
} difference on the pirani gauge. We have done this several times with the same
} results. We are going to try to pressurize the system with He and try He
} sniffer
} to help us maybe pin point the leak. Thanks to all who have responded with
} suggestion on finding the leak and if we are successful in finding the leak I
} will post the finding on the server. Again JEOL has been very responsive in
} trying to find the leak and they are somewhat limited on this instrument do
} to
} its environment. Any other suggestion are welcomed.
}
}
} Bruce W. Arey
}
} PNNL-Battelle
} Associate Scientist
} Microstructural Characterization
} bruce.arey-at-pnl.gov
} 509-376-3363
} fax 509-376-6308
}
}
} ----------
} From: "Dopeyee-at-aol.com"-at-sparc5.microscopy.com
} Sent: Thursday, June 8, 2000 4:28 PM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: LEAK IN JEOL
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} GET A HOLD OF A LEAK DETECTOR AND FIND THE TRUE AREA THAT IS LEAKING
}
} I DO BELEIVE THE MANUFACTURER SHOULD BE OF MORE HELP
}
} THEY HAVE TO FIX IT!!!!
}





From: jim :      jim-at-proscitech.com.au
Date: Sat, 10 Jun 2000 18:59:21 +1000
Subject: RE: Cleaning optical lenses and filters.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hmmm, and "I didn't even know that", but believed that deionised water had
metal ions removed from it and so in that respect its purer than 2x glass
distilled water. Then I was taught and believed! that metal pipes would
re-introduce metal ion back into the water! Being deionised the water has no
buffering capacity and therefore is neither acid nor alkaline, they told me and
I believed.
Education is expensive; must ask for my money back.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Saturday, June 10, 2000 12:24 AM, "milesd-at-us.ibm.com"-at-sparc5.microscopy.com
[SMTP:"milesd-at-us.ibm.com"-at-sparc5.microscopy.com] wrote:
}
}
} I have always understood that deionized water would etch metal,
} that being the reason for PVC pipe being used to deliver it. If this
} is true, wouldn't that damage the first surface mirrors?
}
} Darrell
}





From: milesd-at-us.ibm.com
Date: Sat, 10 Jun 2000 15:24:26 -0400
Subject: RE: Cleaning optical lenses and filters.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ahhh, but I have now been educated! Re-contamination of the
painstakingly purified water is the concern, and there is no
threat to the durability of the pipes. I had been mislead.

Darrell





From: Anaspec :      anaspec-at-icon.co.za
Date: Mon, 12 Jun 2000 07:57:51 +0200
Subject: RE: LEAK IN JEOL

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi
Vacuum leaks, what a pleasure!
Our tried and tested methods include using Petroleum Ether or Ethanol and
then Bostick Prestic or Blue tac as it is known in Australia.
Spraying alcohol around the suspected areas should show up the leak.
If you ant to try and stop a leak use the Prestic. Its that stuff you buy at
the stationary shop that is used to stick posters and pictures to a wall.
That is pliable, removable and really handy at sealing off a few suspect
areas.
We are currently working on a JEOL840 here with a leak. After many hours we
found that it was the gauge head that was leaking.

Good luck
Luc Harmsen
Anaspec, South Africa
Technical support on microscopy.
Tel + 27 (0) 11 476 3455
www.anaspec.co.za

ICEM 15 will be in Durban,
South Africa, 2002.
www.icem15.com

-----Original Message-----
} From: Arey, Bruce W [mailto:bruce.arey-at-pnl.gov]
Sent: Friday, June 09, 2000 4:05 PM
To: 'microscopy-at-msa.microscopy.com'


JEOL has been very helpful and they are somewhat limited on what they can do
to
this instrument. The 840 has radioactive particles inside the chamber so we
have
to be careful each time we open the system up plus there hands on is limited
to
just oversight. I am in the process of trying to find a leak detector on
site
that can be used on radioactive system. But we also cannot pump down the
chamber
enough to use a leak detector. So we are in the process of making a plate
that
will fit over the door. We have taken off all our accessories off the
chamber
(EDS, OIM, BSE) and have pressurized the system and have found some evidence
of
a leak on the left hand side on the door, we have changed the o-ring and
still
the seal leaks on this side of the door. We have polished the door seal to
make
sure there is no major scratches or marks. We are pretty confident that
specimen
exchange port is OK we can pump this down and we see no signs of a leak. Why
are
we focusing our attention on the large o-ring and chamber door. We can
clean
the o-ring with alcohol (methanol or ethanol) and we can rough pump the
chamber
out and put the chamber on the high vac system but after a period of time
the
alcohol dries out and the system shuts down too a poor vacuum (overnight).
Then
we try to rough pump the chamber and we cannot go more than 20-30 mamps
difference on the pirani gauge. We have done this several times with the
same
results. We are going to try to pressurize the system with He and try He
sniffer
to help us maybe pin point the leak. Thanks to all who have responded with
suggestion on finding the leak and if we are successful in finding the leak
I
will post the finding on the server. Again JEOL has been very responsive in
trying to find the leak and they are somewhat limited on this instrument do
to
its environment. Any other suggestion are welcomed.


Bruce W. Arey

PNNL-Battelle
Associate Scientist
Microstructural Characterization
bruce.arey-at-pnl.gov
509-376-3363
fax 509-376-6308


----------
From: "Dopeyee-at-aol.com"-at-sparc5.microscopy.com
Sent: Thursday, June 8, 2000 4:28 PM
To: Microscopy-at-sparc5.microscopy.com
Subject: LEAK IN JEOL

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America
To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
-----------------------------------------------------------------------.


GET A HOLD OF A LEAK DETECTOR AND FIND THE TRUE AREA THAT IS LEAKING

I DO BELEIVE THE MANUFACTURER SHOULD BE OF MORE HELP

THEY HAVE TO FIX IT!!!!






From: Pauline C. Yu :      splene-at-pw.usda.gov
Date: Mon, 12 Jun 2000 08:29:05 -0700 (PDT)
Subject: SEM of liquid crystal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can anyone recommend a book or preferably a website with SEM images of
liquid crystal and or liquid crystal displays?

TX
Pauline Yu
pyu-at-pw.usda.gov
Microscopist Technician
USDA-ARS-WRRC





From: Norman Michaud :      Norman_Michaud-at-meei.harvard.edu
Date: Mon, 12 Jun 2000 16:09:40 -0400
Subject: TEM lab equipment available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are consolidating two electron microscopy labs in the Boston area and
have some equipment which someone may want.
1) JEOL JEM 100CX electron microscope which has always been under service
contract and is still being used. It is about 20 years old and we would like
to see it in a new home rather than trashing it. You would need to have it
moved.
2)JEOL JEE 4C vacuum evaporator for Carbon. Still under vacuum and yours for
the taking.
3)Durst Laborator 138S floor model enlarger with many condensers and an Agfa
Rapidoprint DD6400 processor, both in very good shape and for sale.
Contact me directly with any questions.


Norman Michaud
Director, Morphology
Mass Eye and Ear Infirmary
Ophthalmology-5th flr.
243 Charles St, Boston, MA 02114
norman_michaud-at-MEEI.Harvard.edu
Tel:617-573-3316; Fax:617-573-4290





From: Hans Brinkies :      HBrinkies-at-groupwise.swin.edu.au
Date: Tue, 13 Jun 2000 08:48:34 +1000
Subject: Radioactice SEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bruce W. Arey referred in his request to find a vacuum leak in his
JSM840 to 'radioactive particles' inside the chamber of his SEM. I
wonder if he could be more specific, because I feel if we start calling
a SEM as a 'radioactive system' many safety officers will have a field
day. I have been working with various SEMs since 1968 and never felt
that I might be bombarded by radioactive particles, even when I opened
the chamber (e.g. ETEC) of the SEM.



Hans Brinkies
Senior Lecturer
Swinburne, University of Technology
School of Engineering and Science
P.O.Box 218 - Hawthorn - Vic -3122 - Australia
Phone: +61 3 9214 8657
Fax: +61 3 9214 8264
Email: Hbrinkies-at-swin.edu.au





From: Gillian Bond :      gbond-at-nmt.edu
Date: Mon, 12 Jun 2000 17:43:23 -0600 (MDT)
Subject: Jet polishing of tungsten

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have a student who has been trying to jet polish tungsten for TEM. She
has tried various concentrations of sodium hydroxide in water, as well as
40g trisodium phosphate/250ml water, and 55.8g magnesium perchlorate/250ml
methanol, at a range of voltages. We have a Fischione jet-polishing
unit. So far, none of the samples has been close to good. Can anyone
help us out here, with past experience or general suggestions?

Many thanks in advance,

Gill

Dr Gillian M. Bond
Department of Materials & Metallurgical Engineering
New Mexico Tech
Socorro, NM 87801





From: David Henriks :      Henriks-at-CompuServe.COM
Date: Mon, 12 Jun 2000 22:39:40 -0400
Subject: Jet polishing of tungsten

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Gill:

Certainly the best reference you can have for any jet polishing inquiry is
Bernie Kestel at Argonne National Laboratory. I will forward this to him
to see if he can add anything else. In digging through my extensive
"Bernie Archives" I did find a paper titled "A Jet Polishing Solution for
Silicon Germanium, Tantalum, Niobium and Tungsten-Rhenium" Ultramicrscopy 9
(1982) 379-384.

He was able to get a good polish under the following conditions:
Temperature: -50 C
Jet Height: 4.5mm (Single vertical jet system)
Pump setting: 6
Volts: 40
Current: 20mA

This was done using his BK-1 solution. BK-1 is prepared by mixing 500ml
methanol, 100ml butyl cellosolve, 90ml H2SO4 and 30ml HF.

He also has another paper MRS Volume 199 "Improved Methods and Novel
Techniques for Jet Electropolishing of TEM Foils" which lists a method for
electropolishing a 0.010" tungsten wire.

He was able to get a good polish under the following conditions:
Temperature: -50 C
Pump setting: 6
Volts: 120V

Using 6% HF, 12% sulphuric acid, 68% methanol, and 14% butyl cellosolve.

Of course these were done with a South Bay Vertical Jet system so you will
need to adjust the parameters for your system. Please get the reference
papers or contact me and I will send them to you. I have a long list of
Bernie's papers I could send you along with many other references on TEM
sample preparation that may be of interest. If you'd like to see more,
please contact me.

DISCLAIMER: South Bay Technology produces the Model 550 D Single Vertical
Jet Electropolisher as described above and, therefore, has a vested
interest in promoting its use.

Best regards-

David
Writing at 7:27:57 PM on 06/12/2000

***************************************************************************
************************

David Henriks
Vice President TEL: 800-728-2233
(toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499
San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com

***************************************************************************
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.

Message text written by Gillian Bond
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America



I have a student who has been trying to jet polish tungsten for TEM. She
has tried various concentrations of sodium hydroxide in water, as well as
40g trisodium phosphate/250ml water, and 55.8g magnesium perchlorate/250ml
methanol, at a range of voltages. We have a Fischione jet-polishing
unit. So far, none of the samples has been close to good. Can anyone
help us out here, with past experience or general suggestions?

Many thanks in advance,

Gill

Dr Gillian M. Bond
Department of Materials & Metallurgical Engineering
New Mexico Tech
Socorro, NM 87801
{




From: Tommy Davis :      barb29t-at-arabia.com
Date: Mon, 12 Jun 2000 22:42:05 -0500
Subject: Disaster Recovery Planning #5453

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Business continuity and disaster recovery planning has now
been made easier than ever with Version 7.3 of the
Disaster Recovery System (DRS) product. DRS provides a
plan for inaccessibility or inoperability (a disaster situation).

DRS is an industry standard software product, used by thousands
worldwide. DRS users are in large and small companies across
a wide variety of industries.

DRS conforms to federal regulations and meets insurance, auditing
and legal requirements. DRS runs under Windows, with stand-alone
and network versions available. DRS provides the most complete,
easy-to-use product available today.

To prove its value, a free trial is available. For more
information, visit our web site at
www.drsbytamp.com {http://www.drsbytamp.com} .
Dealer and distributor inquiries are welcomed

*****************************************************************
This message is sent in compliance of the new email bill
section 301. Per Section 301, Paragraph (a)(2)(C) of S. 1618,
further transmissions to you by the sender of this email may be
stopped at no cost to you. This message is not intended for
residents in the State of WA, CA & VA Screening of addresses
has been done to the best of our technical ability.If you are a
Washington, Virginia, or California resident please remove
yourself.
====================================================

///////////////////////////////////////////////////
To be removed permanantly from this list reply to:
mailto:bkf21b-at-netscape.net?subject=remove
///////////////////////////////////////////////////







From: Tommy Davis :      barb29t-at-arabia.com
Date: Mon, 12 Jun 2000 22:42:05 -0500
Subject: Disaster Recovery Planning #5453

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Business continuity and disaster recovery planning has now
been made easier than ever with Version 7.3 of the
Disaster Recovery System (DRS) product. DRS provides a
plan for inaccessibility or inoperability (a disaster situation).

DRS is an industry standard software product, used by thousands
worldwide. DRS users are in large and small companies across
a wide variety of industries.

DRS conforms to federal regulations and meets insurance, auditing
and legal requirements. DRS runs under Windows, with stand-alone
and network versions available. DRS provides the most complete,
easy-to-use product available today.

To prove its value, a free trial is available. For more
information, visit our web site at
www.drsbytamp.com {http://www.drsbytamp.com} .
Dealer and distributor inquiries are welcomed

*****************************************************************
This message is sent in compliance of the new email bill
section 301. Per Section 301, Paragraph (a)(2)(C) of S. 1618,
further transmissions to you by the sender of this email may be
stopped at no cost to you. This message is not intended for
residents in the State of WA, CA & VA Screening of addresses
has been done to the best of our technical ability.If you are a
Washington, Virginia, or California resident please remove
yourself.
====================================================

///////////////////////////////////////////////////
To be removed permanantly from this list reply to:
mailto:bkf21b-at-netscape.net?subject=remove
///////////////////////////////////////////////////







From: Per Hšrstedt :      per.horstedt-at-pathol.umu.se
Date: Tue, 13 Jun 2000 09:17:46 +0200
Subject: Zeiss EM 109 vacuum question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
one of our TEM's , a Zeiss EM 109, is equipped with an ion getter hi-vac
pump. We are now planning to install a turbopump including all the piping
and valves that are necessary. The reason for this is that the high vacuum
is not good enough ( only in low E-6 area ) and the ion pump has to work
too hard and the life-time of the electrodes becomes very short. Also the
housing is clogged with trapped gas molecules and has to be regenerated far
too often.
Technically and electronically this exchange is fairly easily done, but my
question is if anyone has done it and if so, what's the experience?
Yours sincerely

Per Hšrstedt
Department of Medical Biosciences
Pathology
Unit for Electron Microscopy
University of UmeŚ
S-90187 UmeŚ
Sweden

phone int-46-90-7851541
fax int-46-90-7851215





From: Alexander Mironov :      mironov-at-cmns.mnegri.it
Date: Tue, 13 Jun 2000 11:31:30 +0200 (MET DST)
Subject: About monovalent Fab fragments conjugated with HRP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friends,
For a long time we used Fab fragments-HRP from BioSys (France) for the
immuno EM labelling of proteins with subsequent preembedding. These
fragments were the best. However, recently the company cancelled its
activity and does not send any more these products.
Would you be so kind to tell me what has happened (if you know) and what
fragments have the same quality?

Sincerely yours, Alexander Mironov





From: White, Woody N :      Woody.N.White-at-mcdermott.com
Date: Tue, 13 Jun 2000 08:05:00 -0500
Subject: Re:Zeiss EM 109 vacuum question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I changed my former SEM (Etec w/D.P)to a a Leybold mag-lev turbo a number of
years ago. The ball bearing pumps (of the day) caused too much vibration.
The
results were excellent! The Etec plumbing allowed the turbo to run
continuously
during vent cycles which was of great benefit.

Keep in mind that for any gas, turbos have a fixed compression ratio. Among
other things, the foreline pressure will have a direct influence on the
ultimate
vacuum.

The ion pumos are better than turbos for the cleanest, highest vacuum, but
you
have observed one weakness. They do not excel at pumping large volumes of
garbage-loaded gasses.

Good Luck,

Woody White
McDermott Technology

Hi,
one of our TEM's , a Zeiss EM 109, is equipped with an ion getter hi-vac
pump. We are now planning to install a turbopump including all the piping
SNIP




From: White, Woody N :      Woody.N.White-at-mcdermott.com
Date: Tue, 13 Jun 2000 08:10:00 -0500
Subject: Re:Radioactice SEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I believe The SEM to which Bruce refered was/is used to examine radioactive
materials. The "zoomies" are not from the SEM itself, but contamination
from
his specimens. My (former) Etec was in a similar condition. Over the
years, my
work mix resulted in a stage/chamber activitly level in the thousands of cpm
generated by radioactive products of nuclear fission. ...Kept the really
loose
stuff at a minimum, but certainly had to exercise the appropriate
precautions
when working on the system.

Woody White
McDermott Technology
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Bruce W. Arey referred in his request to find a vacuum leak in his
JSM840 to 'radioactive particles' inside the chamber of his SEM. I
wonder if he could be more specific, because I feel if we start calling
a SEM as a 'radioactive system' many safety officers will have a field
day. I have been working with various SEMs since 1968 and never felt
that I might be bombarded by radioactive particles, even when I opened
the chamber (e.g. ETEC) of the SEM.



Hans Brinkies
Senior Lecturer
Swinburne, University of Technology
School of Engineering and Science
P.O.Box 218 - Hawthorn - Vic -3122 - Australia
Phone: +61 3 9214 8657
Fax: +61 3 9214 8264
Email: Hbrinkies-at-swin.edu.au




From: Alexander Mironov :      mironov-at-cmns.mnegri.it
Date: Tue, 13 Jun 2000 07:33:27 -0500
Subject: TEM. About Fab-HRP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friends,
For a long time we used Fab fragments-HRP from BioSys (France) for the
immuno EM labelling of proteins with subsequent preembedding. These
fragments were the best. However, recently the company cancelled its
activity and does not send any more these products.
Would you be so kind to tell me what has happened (if you know) and what
fragments have the same quality?

Sincerely yours, Alexander Mironov






From: LAB. DE MICROSCOPIA ELECTRONICA - FI - UNER :      microsc-at-fi.uner.edu.ar
Date: Tue, 13 Jun 2000 09:50:51 +0000
Subject: looking for a critical point drier...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Laboratory of Electron Microscopy is looking for an used critical point
drier to obtain it in donation....
We can pay all the costs of shipping and handling
For any questions, please
contact to me....

best regards....
===================================================
Fernando D. Balducci
Laboratorio de Microscopia Electr—nica
Facultad de Ingenier’a - Bioingenieria
Universidad Nacional de Entre Rios
Argentina
e-mail: microsc-at-fi.uner.edu.ar
tel: 54 43 975100
fax: 54 43 975077


===================================================




From: Ron Doole :      ron.doole-at-materials.oxford.ac.uk
Date: Tue, 13 Jun 2000 14:08:45 +0100 (GMT Daylight Time)
Subject: Re: Zeiss EM 109 vacuum question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Per,

I have done this on a JEOL 4000 but not a Zeiss. I used a
Maglev TMP (Seiko Seiki) and an antivibration bellows to
couple the pump because I was afraid of vibration degrading
the 0.25nm resolution. I need not have worried, even with
the antivibration bellows shorted out I was OK.
Check that the TMP you choose does not give out any
magnetic fields when running.
If vibration is a problem then the Balzers (Pfieiffer)
antivibration bellows that has a large worm drive clip
around them can be tuned (by tightening the clip) to avoid
the instrument resonant frequency.

Good luck,
Ron

----------------------
Mr. R.C. Doole
Department of Materials,
University of Oxford.
Parks Road, Oxford. OX1 3PH. UK.
Phone +44 (0) 1865 273701
Fax +44 (0) 1865 283333
ron.doole-at-materials.ox.ac.uk

On Tue, 13 Jun 2000 09:17:46 +0200 Per
=?iso-8859-1?Q?H=F6rstedt?= {per.horstedt-at-pathol.umu.se}
wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi,
} one of our TEM's , a Zeiss EM 109, is equipped with an ion getter hi-vac
} pump. We are now planning to install a turbopump including all the piping
} and valves that are necessary. The reason for this is that the high vacuum
} is not good enough ( only in low E-6 area ) and the ion pump has to work
} too hard and the life-time of the electrodes becomes very short. Also the
} housing is clogged with trapped gas molecules and has to be regenerated far
} too often.
} Technically and electronically this exchange is fairly easily done, but my
} question is if anyone has done it and if so, what's the experience?
} Yours sincerely
}
} Per Hörstedt
} Department of Medical Biosciences
} Pathology
} Unit for Electron Microscopy
} University of Umeĺ
} S-90187 Umeĺ
} Sweden
}
} phone int-46-90-7851541
} fax int-46-90-7851215





From: Phoebe J Doss :      pjdoss-at-okstate.edu
Date: Tue, 13 Jun 2000 09:32:35 -0500
Subject: help, I've fallen and can't get rid of the precipitate!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Hello:

After many years of staining grids with uranyl acetate and lead citrate, we
have begun to see a needle like or shard precipitate, (about 1/2 inch long
at 100,000x's; resembles the lead precipitate on page 469 of Electron
Microscopy second edition John Bozzola and Lonnie Russell). We have been
using a 2.5% aqueous uranyl acetate and Reynold's lead citrate (filtered
through a 2 micron filter) for the past 4 years or so with no problems. I
have stained the grids with just UA and can see no precipitate and have
stained grids with just the lead citrate and still not see the precipitate.
I have also checked the water and cannot still see the precipitate.
However, when I stain with UA followed by lead citrate it mysteriously
reappears much to my dissatisfaction. I have also tried the basic lead
citrate and just recently tried Sato's lead stain and had the same problem.
I have made up UA from a newly purchased bottle. I have also lessened the
staining times from 30 minutes in UA to 7 minutes and from 20 minutes in
lead citrate to 5 minutes and the precipitate is less but still there. I
have also checked the grids before staining them and cannot see the
precipitate. Please help, I grow more grey day by day.

Phoebe J. Doss
Manager/Adjunct Instructor
Electron Microscope Lab
Oklahoma State University





From: Tamara Howard :      howard-at-cshl.org
Date: Tue, 13 Jun 2000 10:31:29 -0400 (EDT)
Subject: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Back in the olden days, when BioRad sold microscopy supplies, they had an
epoxy cleaner (to remove epoxies from hands, benches, etc., not made from
epoxies). It was blue gunk in a little jar. Does anyone know what happened
to this stuff? Or have an alternate?

I've just been digging through catalogues and Ted Pella sells a liquid
cleaner - any experience with it?

Thanks!

Tamara Howard
CSHL





From: Ronald LHerault :      lherault-at-bu.edu
Date: Tue, 13 Jun 2000 10:54:11 -0400 (EDT)
Subject: Streptococci Mutans preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


a student here needs to fix and dehydrate strep mutans on polystyrene
petri dishes. Can anyone point me to a good (simple?) protocol? Since we
are primarily a materials research lab, we don't have a lot of biological
references.

Thanks

Ron L
lherault-at-bu.edu





From: Beverly_E_Maleeff-at-sbphrd.com
Date: Tue, 13 Jun 2000 12:15:16 -0400
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tamara:

I don't know the Bio-Rad product, but Ladd sold (still sells?) a product called
Met-a-terge that gets rid of uncured resins. A little bit goes a long way.
I've used it for years.

Hope this helps.

Regards,
Bev Maleeff
SmithKline Beecham Pharmaceuticals






From: Anthony Garratt-Reed :      tonygr-at-mit.edu
Date: Tue, 13 Jun 2000 13:08:35 -0400
Subject: Transfer of University-owned instrumentation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Mansfield's post regarding surplus equipment prompts me to make the
following observations regarding the transfer of University-owned equipment
in the United States. I would imagine that many other countries might have
similar policies.

It often seems to come as a surprise to people to discover that the US
government won't buy the same piece of equipment twice. What I mean by
this is that if a piece of equipment has originally been purchased using
federal funds (regardless of who currently holds title), then another
institution cannot use federal funds (regardless of the source) to buy the
equipment from the first owner.

To use the specific example of John's equipment: suppose it was bought
originally with, say, an NSF grant, and I find that I could make use of it
now. In order to do that I would need to find a non-US-government source
of money, as I could not use even the income from my facility operation
(which is regarded by the accountants as government money, as it originates
predominantly from government research grants).

The logic of this policy, of course, is quite inescapable, however
unpalatable it may be to the present owners of the equipment.

The policy only covers the cost of purchasing the equipment, by the way.
If, to use the same example, John were to give me his surplus equipment, it
would be perfectly acceptable for me to use federal funds to pay for the
packing and shipping (and, if appropriate, reinstallation)

Tony Garratt-Reed.



** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6479
**






From: drose-at-wlgore.com
Date: Tue, 13 Jun 2000 13:44:37 -0400
Subject: SEM - sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Dear List,

I am using a Denton Desk II sputter coater with a Pt target. I have noticed
recently that a dark spot has formed in the middle of the target. I have not
seen this before. Is this due to impurities in the target, problems with the
vacuum, contamination from outgassing samples, or something else? Everything
appears to be running fine and sample types have not changed. Do I clean it or
just leave it alone? What is it telling me (if the coater could talk)?

Thanks in advance for all your expertise.

David BG Rose
WL Gore and Associates
297 BLue Ball Road
Elkton, MD 21921
410-506-2958






From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Tue, 13 Jun 2000 13:13:50 -0500
Subject: help, I've fallen and can't get rid of the precipitate!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Phoebe,

We recently had a rash of horrible precipitate problems that we couldn't
seem to trace. Our staining procedures are similar to yours. We made up
fresh stains, changed all our syringe filters, used every precaution we
could think of.

Then we had the water system checked. Our in-line reverse-osmosis,
deionization system had become a mess, although we had assumed (there's that
word!) that the company we leased it from was maintaining it properly.
Turns out that they thought we owned the system, while in fact we only
rented it and paid them to service it.

Anyway, to keep it short, we purchased a Millipore bench-top, low-volume
water-polishing unit and used our old water to feed it (after getting the
thing serviced properly). Our precipitates disappeared and have not yet
reappeared.

For what it's worth.

(No financial interest in Millipore, etc.)

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/




-----Original Message-----
} From: Phoebe J Doss [mailto:pjdoss-at-okstate.edu]
Sent: Tuesday, June 13, 2000 9:33 AM
To: microscopy-at-sparc5.microscopy.com


Hello:

After many years of staining grids with uranyl acetate and lead citrate, we
have begun to see a needle like or shard precipitate, (about 1/2 inch long
at 100,000x's; resembles the lead precipitate on page 469 of Electron
Microscopy second edition John Bozzola and Lonnie Russell). We have been
using a 2.5% aqueous uranyl acetate and Reynold's lead citrate (filtered
through a 2 micron filter) for the past 4 years or so with no problems. I
have stained the grids with just UA and can see no precipitate and have
stained grids with just the lead citrate and still not see the precipitate.
I have also checked the water and cannot still see the precipitate.
However, when I stain with UA followed by lead citrate it mysteriously
reappears much to my dissatisfaction. I have also tried the basic lead
citrate and just recently tried Sato's lead stain and had the same problem.
I have made up UA from a newly purchased bottle. I have also lessened the
staining times from 30 minutes in UA to 7 minutes and from 20 minutes in
lead citrate to 5 minutes and the precipitate is less but still there. I
have also checked the grids before staining them and cannot see the
precipitate. Please help, I grow more grey day by day.

Phoebe J. Doss
Manager/Adjunct Instructor
Electron Microscope Lab
Oklahoma State University





From: Snow, David B. :      snowdb-at-pweh.com
Date: Tue, 13 Jun 2000 13:17:57 -0500
Subject: Transmission Electron Microscopist (TEM) / Engineer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Transmission Electron Microscopist (TEM) / Engineer

United Technologies Corporation is seeking an engineer to fill the TEM
operator/engineer position at the United Technologies Research Center in
East Hartford, CT. This position will provide support to the United
Technologies Corporation Business Units including Pratt & Whitney, Carrier,
Sikorsky Aircraft, Hamilton Sundstrand, Otis Elevator, and International
Fuel Cells. The TEM operator will be responsible for the dailyoperations of
the TEM laboratory; including preparation of TEM samples using various
techniques such as dimpling, ion milling, jet polishing, microtoming, and
replication. Project duties include conducting failure analyses,
characterization of surface coatings, and analysis of advanced metal and
ceramic materials. The candidate should have experience with both TEM
sample preparation and conventional TEM operation. Good communication and
interpersonal skills are essential. The ability to recognize fracture modes
and origins of fractures is desired. Experience with Scanning Electron
Microscopy (SEM) is a plus.

Qualified candidates will have a BS in Materials Science or equivalent, with
a minimum of 2 years TEM experience. U. S. citizenship or permanent
residency is required.

Please visit our web site at www.utrc.utc.com, and send your resume to
Employment Opportunities, Code MATS-2050-9049, 411 Silver Lane, East
Hartford, CT 06118 or e-mail employment-at-utrc.utc.com. United Technologies
Corporation is an equal opportunity employer.







From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Wed, 14 Jun 2000 08:27:16 GMT+1200
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, Tamara

}
} Back in the olden days, when BioRad sold microscopy supplies, they had an
} epoxy cleaner (to remove epoxies from hands, benches, etc., not made from
} epoxies). It was blue gunk in a little jar. Does anyone know what happened
} to this stuff? Or have an alternate?
}
} I've just been digging through catalogues and Ted Pella sells a liquid
} cleaner - any experience with it?
}

I seem to remember that soaking in N,N-dimethyl formamide dissolves
epoxy, you might care to beg a little from a freiendly chemistry dept
and try it, it's not very expensive. It has a moderately offensive
smeel, though.

rtch


Ritchie Sims Phone : 64 9 3737599 ext 7713
Department of Geology Fax : 64 9 3737435
The University of Auckland email : r.sims-at-auckland.ac.nz
Private Bag 92019
Auckland
New Zealand




From: Ladd Research :      ladres-at-worldnet.att.net
Date: Tue, 13 Jun 2000 16:35:42 -0400
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tamara, Beverly and all other interested parties,

Yes, Ladd stills sells Met-A-Terge (catalog #13045).
Please check our web site http://www.laddresearch.com
for more information or contact me off line.

JD Arnott

Disclaimer: As stated above, Ladd sell Met-A-Terge and thus has a
commercial interest in this thread.

LADD RESEARCH
131 Dorset Lane
Williston, VT 05495 USA

TEL 1-800-451-3406 (US) or 1-802-878-6711 (anywhere)
FAX 1-802-878-8074
e-mail ladres-at-worldnet.att.net

ALL NEW WEB SITE UP NOW AT OUR NEW URL:

http://www.laddresearch.com


Beverly_E_Maleeff-at-sbphrd.com-at-sparc5.microscopy.com wrote:
}
} ------------------------------------------------------------------------} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.}
} Tamara:
}
} I don't know the Bio-Rad product, but Ladd sold (still sells?) a product called
} Met-a-terge that gets rid of uncured resins. A little bit goes a long way.
} I've used it for years.
}
} Hope this helps.
}
} Regards,
} Bev Maleeff
} SmithKline Beecham Pharmaceuticals

--




From: Jo Dee Fish :      jofish-at-burnham-inst.org
Date: Tue, 13 Jun 2000 13:38:18 -0700
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Tamara,
I have used the Ted Pella Epoxy Hand Cleaner and it works well. It will
remove epoxy from hands and also glassware.
Jo Dee Fish

PS I am not affiliated with Ted Pella, just love their products!

Tamara Howard wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Back in the olden days, when BioRad sold microscopy supplies, they had an
} epoxy cleaner (to remove epoxies from hands, benches, etc., not made from
} epoxies). It was blue gunk in a little jar. Does anyone know what happened
} to this stuff? Or have an alternate?
}
} I've just been digging through catalogues and Ted Pella sells a liquid
} cleaner - any experience with it?
}
} Thanks!
}
} Tamara Howard
} CSHL

--
Jo Dee Fish
Electron Microscopy Assistant
Cell Analysis Facility
The Burnham Institute
10901 N. Torrey Pines Rd.
La Jolla, CA 92037
(858)646-3100 ext. 3620






From: Connie McManus :      conmac-at-cc.usu.edu
Date: Tue, 13 Jun 2000 15:29:58 -0600
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My husband uses epoxy on a sailboat he's building. He cleans everything
up... hands, spills, etc... with plain old vinegar. We go through a LOT of
vinegar. If it's dried, then soak acetone on it until it softens, then use
vinegar (or 5% acetic acid) to mop up the residues.

connie m

At 10:31 AM 06/13/2000 -0400, Tamara Howard wrote:
} Back in the olden days, when BioRad sold microscopy supplies, they had an
} epoxy cleaner (to remove epoxies from hands, benches, etc., not made from
} epoxies). It was blue gunk in a little jar. Does anyone know what happened
} to this stuff? Or have an alternate?
}
} I've just been digging through catalogues and Ted Pella sells a liquid
} cleaner - any experience with it?
}
} Thanks!
}
} Tamara Howard
} CSHL
}
}
}
Connie McManus
Veterinary Diagnostics Lab
Utah State University
Logan, UT
USA




From: =shAf= :      mshaf-at-darkwing.uoregon.edu
Date: Tue, 13 Jun 2000 16:26:55 -0700
Subject: RE: Transfer of University-owned instrumentation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Anthony Garratt-Reed writes ...

} To use the specific example of John's equipment: suppose
} it was bought originally with, say, an NSF grant, and
} I find that I could make use of it now. In order to do
} that I would need to find a non-US-government source
} of money, as I could not use even the income from my
} facility operation (which is regarded by the accountants
} as government money, as it originates
} predominantly from government research grants).

I would assume it can even get messy if, at least some, of my
facility's income had come from outside sources. I would imagine the
accountants will first assume it is government $$ ... in which case I
would have to show I had taken in an equivelent in outside $$ ... but
over what time frame? ... this fiscal? ... last 10 years?

cheerios, =shAf= :o)

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - mshaf-at-darkwing.uoregon.edu
Geological Science's Electron Probe Facility - University of Oregon
http://epmalab.uoregon.edu/






From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Tue, 13 Jun 2000 19:51:28 -0700
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I use methylene dichloride (dimethyl chloride) to dissolve most any epoxy.

I have 3-5 gallons of it. I do not know if it is still generally available.

gg

At 07:31 AM 6/13/00, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: earlw-at-pacbell.net
Date: Tue, 13 Jun 2000 20:43:51 -0700
Subject: Re: MSA Listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Joe,

The listserver is at MSA listserver {Microscopy-at-sparc5.microscopy.com}
Just sent email say, "Please subscribe".
It is fairly active but is only about 1/2 relevant as most data is biology
oriented.

Earl
JCNABITY-at-aol.com wrote:

} Dear Earl,
}
} Could you tell me how to subscribe to the MSA listserver? Greg talked highly
} of it and I'm thinking I will set up another e-mail account to use for it.
} Since it is pretty active, I didn't want all the messages going in with my
} normal e-mail, but a using a separate account will resolve that issue.
}
} Joe





From: mohamed a jaffer :      mohamed-at-molbiol.uct.ac.za
Date: Wed, 14 Jun 2000 13:45:54 SAST-2
Subject: 70mm roll film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear List

Our EM Unit has a Zeiss EM109 which uses 70mm roll film - Agfa
Scientia 23D56. Manufacture of this film ceased quite a while back.
We are trying to locate unused stock of this film for our usage. If
you have any surplus stock for sale please contact me.

Thanks
Mohamed

******************************
M. A. Jaffer
Electron Microscope Unit
R. W. James Building
University of Cape Town
Private Bag
Rondebosch, 7701
South Africa

Tel: +27-21-6503354
fax: +27-21-6891528

e-mail: mohamed-at-molbiol.uct.ac.za
***************************************************




From: Connie McManus :      conmac-at-cc.usu.edu
Date: Wed, 14 Jun 2000 07:47:42 -0600
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 06:09 PM 06/13/2000 -0700, Don Hammer wrote:
} Stuff is cheap too and if there is any left after the zillions of home uses,
} great on salads!!!!

yeah, especially the used stuff........ eeeeuewwwwwwww! *G*

connie m
}
} Don Hammer, Retired Guy
} ----- Original Message -----
} From: Connie McManus {conmac-at-cc.usu.edu}
} To: Tamara Howard {howard-at-cshl.org} ; Microscopy Listserver
} {Microscopy-at-sparc5.microscopy.com} ; Histology listserver
} {histonet-at-pathology.swmed.edu}
} Sent: Tuesday, June 13, 2000 2:29 PM
} Subject: Re: Epoxy cleaner?
}
}
} } My husband uses epoxy on a sailboat he's building. He cleans everything
} } up... hands, spills, etc... with plain old vinegar. We go through a LOT
} of
} } vinegar. If it's dried, then soak acetone on it until it softens, then
} use
} } vinegar (or 5% acetic acid) to mop up the residues.
} }
} } connie m
} }
} } At 10:31 AM 06/13/2000 -0400, Tamara Howard wrote:
} } } Back in the olden days, when BioRad sold microscopy supplies, they had an
} } } epoxy cleaner (to remove epoxies from hands, benches, etc., not made from
} } } epoxies). It was blue gunk in a little jar. Does anyone know what
} happened
} } } to this stuff? Or have an alternate?
} } }
} } } I've just been digging through catalogues and Ted Pella sells a liquid
} } } cleaner - any experience with it?
} } }
} } } Thanks!
} } }
} } } Tamara Howard
} } } CSHL
} } }
} } }
} } }
} } Connie McManus
} } Veterinary Diagnostics Lab
} } Utah State University
} } Logan, UT
} } USA
} }
} }
}
}
Connie McManus
Veterinary Diagnostics Lab
Utah State University
Logan, UT
USA




From: veys-at-bota.ucl.ac.be (Pascal Veys)
Date: Wed, 14 Jun 2000 16:18:08 +0100
Subject: wax and lipid droplets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi to all,
Does anybody have some information on a possible relationship between lipid
droplets and wax production in plant cells. Are there some evidences that
lipid droplets could actually be storage sites of wax precursors ? I
looked for that in literature but found nothing...
Thus : HELP !
References will be welcome
Thanks in advance for answering
Bye
Pascal

""""""""'
( O)(o )
--------------------0000--------------0000----------
Pascal VEYS
Laboratory of Plant Biology
Catholic University of Louvain
Place Croix du Sud 5 (bte 14)
B 1348 Louvain-la-Neuve
Belgium
Phone : 0032 10473004
Fax : 0032 10473471
Email : Veys-at-bota.ucl.ac.be
---------------------ooooO-----------Ooooo--------
( ) (_)(_) ( )
) ( ) (
(_) (_)






From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Wed, 14 Jun 2000 11:40:12 -0500
Subject: Epoxy removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Gary Gaugler wrote:
===============================================================
I use methylene dichloride (dimethyl chloride) to dissolve most any epoxy.
I have 3-5 gallons of it. I do not know if it is still generally available.
================================================================
If we are talking about methylene dichloride, or dichloromethane, CAS # 75-
09-2, this is a pretty bad actor, and is on the list of Prop. 65 chemicals
for the State of California as being cancer causing. Chemicals on the Prop
. 65 list are so highly restricted that in some organizations, they are
allowed in only with the approval of top management.

But I do have a question: I always thought that most epoxies, certainly the
ones used in microscopy, ended up being three dimensionally crosslinked
intractable solids. The only way such a material is going to be "dissolved"
is for chemical bonds to be broken. And yet, I don't see how chemically,
methylene dichloride is going to be breaking chemical bonds. Or the same
comment for some of the other materials mentioned. These materials might
plasticize (e.g. soften) an epoxy and aid in its removal from a surface, but
do any of these really "dissolve" a three dimensionally crosslinked epoxy
system?

I am very interested in this topic because we believe that at least in terms
of getting epoxy out of the "nooks and crannies" of a non-smooth surface, an
oxygen plasma is needed. But perhaps we are wrong about that, that is why
I ask the question.

Disclaimer: SPI Supplies manufactures the Plasma Prep™ II plasma etcher and
has an interest in seeing more applications for plasma etching.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
============================================










From: NPGSlithography-at-aol.com
Date: Wed, 14 Jun 2000 13:28:54 EDT
Subject: Please Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please Subscribe




From: Joyce Craig :      rcraig8-at-crown.icongrp.com
Date: Wed, 14 Jun 2000 13:50:24 -0500
Subject: SEM-sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


David, I had the same problem with a gold target-a dark spot in the
middle. Also, what was being sputtered on my samples was not gold. I
cleaned the target with acetone and it has been working fine since.
Joyce Craig
Chicago State University





From: Marsha Pilgrim :      mpilgrim-at-MendelBio.COM
Date: Wed, 14 Jun 2000 11:51:50 -0700
Subject: uranyl acetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

Can anyone recommend a supplier for uranyl acetate?

Also, does anyone out there have experience with Sigma's Lowicryl kit?

If possible, please respond to me directly at: mpilgrim-at-mendelbio.com

Many thanks,
Marsha





From: Joyce Craig :      rcraig8-at-crown.icongrp.com
Date: Wed, 14 Jun 2000 14:41:55 -0500
Subject: TEM cell cultures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are working with Leishmania major, a parasite that is incorporated
into macrophages. We have had good success with fixation of lymph nodes
infected with Leishmania.
We have not been as happy with the results of fixation of cell cultures
of infected bone marrow macrophages. The membranes of the Leishmania
and of the internal compartments within the macrophages that hold the
Leishmania are well fixed, but the external cell membranes of the
macrophages are somewhat discontiuous. We have not done cell cultures
before. Is this a problem to be expected? We are fixing with 2+2
glutaraldehyde/paraformaldehyde in 0.1 M phosphate buffer, post-fixing
with 2% buffered Osmium, dehydrating with ethanol and propylene oxide,
then embedding in epoxy. We have shortened all times compared to those
we use with tissue samples.
Joyce Craig
Chicago State University





From: Subramanian Karthikeyan :      karthik-at-micro.uoguelph.ca
Date: Wed, 14 Jun 2000 15:38:48 -0500
Subject: non-florescent plastic coverslip..

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I am looking for a non-fluorescent plastic coverslip that allows
confocal laser microscopy and subsequent sectioning for
Transmission Electron Microscopy. I will greatly appreciate your
suggestions.

Sincerely,
Karthi Subramanian
Department of Microbiology
University of Guelph
Guelph Ontario N1G 2W1
Phone: (519)824-4120 ext.8904
Fax:(519)837-1802






From: Dusevich, Vladimir :      DusevichV-at-umkc.edu
Date: Wed, 14 Jun 2000 15:46:01 -0500
Subject: RE: Stereometry by computer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I perform stereo measurements only occasionally,
so I prefer to save money on specialized equipment or
software. All I use is just freeware program ImageTool
(good for on-screen stereo pair measurements) and Excel
(not bad for calculations).

Of course, for a big project it's better to buy a software.

Vladimir M. Dusevich, Ph.D.
Electron Microscope Lab Manager
3127 School of Dentistry
650 E. 25th Street
Kansas City, MO 64108-2784

Phone: (816) 235-2072
Fax: (816) 235-5524
Web: http://www.umkc.edu/dentistry/microscopy

} -----Original Message-----
} From: John J. Bozzola [mailto:bozzola-at-siu.edu]
} Sent: Thursday, June 08, 2000 11:13 AM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: Stereometry by computer
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
} --------------------------------------------------------------
} ---------.
}
}
} A user of our facility is interested in making height measurements
} from specimens viewed in the SEM. I am aware of the conventional way
} of doing this: stereo pairs and optical viewer (with stereometer
} parallax corrections). Is there a more modern (computerized) way of
} doing this, say with anaglyphs?
}
} Thank you.
}
} John B.
}
}
} ####################################################################
} John J. Bozzola, Ph.D., Director
} Micro-Imaging and Analysis Center
} 750 Communications Drive - MC 4402
} Southern Illinois University
} Carbondale, IL 62901 U.S.A.
} Phone: 618-453-3730
} Fax: 618-453-2665
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/departments/shops/cem.html
} ####################################################################
}




From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Wed, 14 Jun 2000 10:48:50 -1000 (HST)
Subject: TEM labeling Si-OH groups?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Oh wise and helpful microscopists-

I need to label Si-OH groups with something that will show up in TEM. If
it were proteins or sugars, I'd immuno- or lectin label. Any ideas how I
can get gold or ferritin or whatever onto these groups?

Mahalo!
Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************





From: Ladd Research :      ladres-at-worldnet.att.net
Date: Wed, 14 Jun 2000 17:03:40 -0400
Subject: Re: uranyl acetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Marsha Pilgrim wrote:

} Hello,
}
} Can anyone recommend a supplier for uranyl acetate?
} Also, does anyone out there have experience with Sigma's Lowicryl kit?
} If possible, please respond to me directly at: mpilgrim-at-mendelbio.com
}
} Many thanks,
} Marsha


Dear Marsha,

We at Ladd Research, and most of the other supply companies, can sell
you this. In our case it is catalog # 23620 and more information can be
found on our web site, http://www.laddresearch.com

John Arnott
--

LADD RESEARCH
131 Dorset Lane
Williston, VT 05495 USA

TEL 1-800-451-3406 (US) or 1-802-878-6711 (anywhere)
FAX 1-802-878-8074
e-mail ladres-at-worldnet.att.net

ALL NEW WEB SITE UP NOW AT OUR NEW URL:

http://www.laddresearch.com




From: beth-at-dogwood.botany.uga.edu (Beth Richardson)
Date: Wed, 14 Jun 2000 17:49:03 -0500
Subject: autofluorescence of plant material

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Help - A faculty member's plant tissue is autofluorescing. She is using
aniline blue to look at pollen tubes (through the styles). She would like
to reduce the background fluorescence. I gave her a copy of a borohyride
reference from a '97 listserv posting. Can anyone recommend a fixation that
will decrease or eliminate the autofluorescence? Any thoughts on this
matter would be greatly appreciated.
thanks,
Beth

**************************************
Beth Richardson
EM Lab Coordinator
Botany Department
University of Georgia
Athens, GA 30602

Phone - (706) 542-1790
FAX - (706) 542-1805
Email - beth-at-dogwood.botany.uga.edu
**************************************






From: Heather A Owen :      owenha-at-csd.uwm.edu
Date: Wed, 14 Jun 2000 16:47:36 -0500 (CDT)
Subject: Re: LR White flat embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


It took me a few days to find this protocol from my collegue Lucinda
Swatzell. It sounded intriguing. Although I've not tried it myself yet,
she's used it with success.

I've copied the following out of her e-mail. Things that need to be kept
in mind are that she polymerizes in a vacuum oven, she's refering to plant
seedlings, and that I don't have any financial interest that I know of
(i.e. I haven't checked my mutual fund prospectus) in Rubbermaid, Inc.

Here it is:

"There is a way to flat embed with LR White: This works great for me when
I am keeping them on their agar blocks and maintaining orientation, but it
should also work for regular seedlings to keep them flat instead of
crooked in the bottom of the capsules. Use rubbermaid ice cube trays.
Place the specimens in the flat bottoms of the tray. cover with about 1/4
in of resin. In each well, place another well that has been cut from it's
tray. The single loose wells will nestle down into the ice cub tray on
top of the resin. Because rubbermaid is dishwasher safe it will take the
heat, but get soft enough to snuggle in tightly and keep out oxygen. When
you pump the vaccuum the extra resin also snakes up into the cracks, so
that you get a seal. Now the thing that is scary: will loose seedlings
just suck up into the cracks? I haven't tried them."

Heather A. Owen, Director
Electron Microscope Laboratory
Department of Biological Sciences
University of Wisconsin - Milwaukee
(414)229-6816







From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Wed, 14 Jun 2000 15:07:34 -0700
Subject: Re: Epoxy removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I've used it often to unpackage epoxy microchips. These
epoxies are silicone/epoxy and not true epoxy. They are
typically called plastic packaged ICs. Whatever. I've
used MEC and DiMEC to do this. Some packages
had to be oxygen ion blasted open. Over time, the plasma
method has been much safer for the operator and the chip.
And it can make the removal process much faster. Plasma
is the current method of choice.

Whether the "epoxies" talked about here are the same as
those for IC packages is likely not to be.

gary g.



At 09:40 AM 6/14/00, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Walck, Scott D. :      walck-at-ppg.com
Date: Wed, 14 Jun 2000 18:26:55 -0400
Subject: RE: Jet polishing of tungsten

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I used to prepare W FIM emitter tips with a 5% NaOH solution and I think the
voltage was around 5 to 10 volts ac. When I used to electropolish, the FIM
solutions and conditions were frequently similar to the jet polishing
solutions.


-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)



} -----Original Message-----
} From: Gillian Bond [mailto:gbond-at-nmt.edu]
} Sent: Monday, June 12, 2000 7:43 PM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: Jet polishing of tungsten
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} --------------------------------------------------------------
} ---------.
}
}
}
} I have a student who has been trying to jet polish tungsten
} for TEM. She
} has tried various concentrations of sodium hydroxide in
} water, as well as
} 40g trisodium phosphate/250ml water, and 55.8g magnesium
} perchlorate/250ml
} methanol, at a range of voltages. We have a Fischione jet-polishing
} unit. So far, none of the samples has been close to good. Can anyone
} help us out here, with past experience or general suggestions?
}
} Many thanks in advance,
}
} Gill
}
} Dr Gillian M. Bond
} Department of Materials & Metallurgical Engineering
} New Mexico Tech
} Socorro, NM 87801
}
}




From: Thomas Neumann :      tneumann-at-u.washington.edu
Date: Wed, 14 Jun 2000 16:43:29 -0700
Subject: Artifact or Structured Water?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My name is Tim Strovas and I am a graduate student at the U of
Washington bioengineering department. I am investigating the halo
effect from single myofibrils (from bumblebee muscle tissue) as seen
under a phase contrast microscope. Has anyone encountered a previous
investigation into this effect and its possible relationship with
structured water?
Note: The Halo does not appear as ripples that are normally associated
with optical microscope artifacts. The halo is single broad band that
borders the tissue sample.






From: Thomas Neumann :      tneumann-at-u.washington.edu
Date: Wed, 14 Jun 2000 16:54:35 -0700
Subject: Artifact or Structured Water?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry, forgot return address: T-Strovas-at-Cornell-Iowa.edu

My name is Tim Strovas and I am a graduate student at the U of
Washington bioengineering department. I am investigating the halo
effect from single myofibrils (from bumblebee muscle tissue) as seen
under a phase contrast microscope. Has anyone encountered a previous
investigation into this effect and its possible relationship with
structured water?
Note: The Halo does not appear as ripples that are normally associated
with optical microscope artifacts. The halo is single broad band that
borders the tissue sample.







From: Thomas Neumann :      tneumann-at-u.washington.edu
Date: Wed, 14 Jun 2000 16:54:23 -0700
Subject: Artifact or Structured Water?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry, forgot return address: T-Strovas-at-Cornell-Iowa.edu

My name is Tim Strovas and I am a graduate student at the U of
Washington bioengineering department. I am investigating the halo
effect from single myofibrils (from bumblebee muscle tissue) as seen
under a phase contrast microscope. Has anyone encountered a previous
investigation into this effect and its possible relationship with
structured water?
Note: The Halo does not appear as ripples that are normally associated
with optical microscope artifacts. The halo is single broad band that
borders the tissue sample.







From: =?iso-2022-jp?B?GyRCI1IjaSN6I2EjbBsoSg==?=
Date: Thu, 15 Jun 2000 09:08:35 +0900
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html







From: John Nailon :      mmjnailo-at-dingo.cc.uq.edu.au
Date: Thu, 15 Jun 2000 11:35:56 +1000
Subject: Re: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


G'day folks,
Be very wary of any Good Epoxy Solvent, do not use it to clean epoxy off
your skin. Anything that is a good solvent for epoxy will probably be a
good solvent for the oils and lipids in/on your skin. These oils and
lipids are your protection against epoxy resins entering your body.
Remember, all epoxy resins at carcinogenic, soap and water are probably
the safest agents to remove epoxies from your skin.
Regards
JVN

Connie McManus wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} At 06:09 PM 06/13/2000 -0700, Don Hammer wrote:
} } Stuff is cheap too and if there is any left after the zillions of home uses,
} } great on salads!!!!
}
} yeah, especially the used stuff........ eeeeuewwwwwwww! *G*
}
} connie m
} }
} } Don Hammer, Retired Guy
} } ----- Original Message -----
} } From: Connie McManus {conmac-at-cc.usu.edu}
} } To: Tamara Howard {howard-at-cshl.org} ; Microscopy Listserver
} } {Microscopy-at-sparc5.microscopy.com} ; Histology listserver
} } {histonet-at-pathology.swmed.edu}
} } Sent: Tuesday, June 13, 2000 2:29 PM
} } Subject: Re: Epoxy cleaner?
} }
} }
} } } My husband uses epoxy on a sailboat he's building. He cleans everything
} } } up... hands, spills, etc... with plain old vinegar. We go through a LOT
} } of
} } } vinegar. If it's dried, then soak acetone on it until it softens, then
} } use
} } } vinegar (or 5% acetic acid) to mop up the residues.
} } }
} } } connie m
} } }
} } } At 10:31 AM 06/13/2000 -0400, Tamara Howard wrote:
} } } } Back in the olden days, when BioRad sold microscopy supplies, they had an
} } } } epoxy cleaner (to remove epoxies from hands, benches, etc., not made from
} } } } epoxies). It was blue gunk in a little jar. Does anyone know what
} } happened
} } } } to this stuff? Or have an alternate?
} } } }
} } } } I've just been digging through catalogues and Ted Pella sells a liquid
} } } } cleaner - any experience with it?
} } } }
} } } } Thanks!
} } } }
} } } } Tamara Howard
} } } } CSHL
} } } }
} } } }
} } } }
} } } Connie McManus
} } } Veterinary Diagnostics Lab
} } } Utah State University
} } } Logan, UT
} } } USA
} } }
} } }
} }
} }
} Connie McManus
} Veterinary Diagnostics Lab
} Utah State University
} Logan, UT
} USA

--
****************************************************
John V Nailon
Operations Manager
Centre for Microscopy and Microanalysis
The University of Queensland
St. Lucia Queensland 4072
Phone: +61-7-3365-4214
Fax: +61-7-3365-4422
WWW: http://www.uq.edu.au/nanoworld/allstaff.html#Nailon
****************************************************




From: Rick Powell at Nanoprobes :      rpowell-at-nanoprobes.com
Date: Thu, 15 Jun 2000 06:38:06 -0500
Subject: Re: TEM labeling Si-OH groups?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I suggest you copy this message to the plant surfaces mail list

To join, send the command
join plant-surfaces firstname lastname
to: mailbase-at-mailbase.ac.uk
"Firstname" can be one or more names or initials. The last
word in this command will be interpreted as the last name.
The email address will be extracted automatically from the
message.

Chris Jeffree

Date sent: Wed, 14 Jun 2000 16:18:08 +0100
To: microscopy-at-sparc5.microscopy.com
} From: veys-at-bota.ucl.ac.be (Pascal Veys)


Hello Tina:

I would approach your problem by either modifying the Si-OH groups with a
hapten and detecting with antibody- or streptavidin-gold, or converting
them to amines or thiols then labeling with a gold labeling reagent
(disclaimer - we make gold labeling reagents). You could introduce amino-
groups at the Si-OH groups using a silylating reagent such as
3-{Tris[2-(2-methoxyethoxy)ethoxy]silyl}propylamine or
3-[Tris(trimethylsiloxy)silyl]propylamine (both from Fluka), then either
biotinylate with NHS-biotin and detect with streptavidin-gold, or label the
amines with Mono-Sulfo-NHS-Nanogold.

I have not actually tried this, and since I don't know what types of
samples you are looking at, it's difficult to say what else in them might
affect the reaction. If there are already other primary amines in your
sample, they need to be blocked first.

If you would like other ideas, a text on solid-phase oligo- or peptide
synthesis might be another good starting point - the chemistry used to
functionalize the beads used in these systems may also be transferable to
your situation.

Hope this is helpful,

Rick Powell


}
} Oh wise and helpful microscopists-
}
} I need to label Si-OH groups with something that will show up in TEM. If
} it were proteins or sugars, I'd immuno- or lectin label. Any ideas how I
} can get gold or ferritin or whatever onto these groups?
}
} Mahalo!
} Tina
}
} ****************************************************************************
} * Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
} * Biological Electron Microscope Facility * (808) 956-6251 *
} * University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
} ****************************************************************************ť

**********************************************************************
* NANOPROBES, Incorporated | Tel: (919) 510-0590 *
* 95 Horse Block Road | Fax: (919) 510-0590 *
* Yaphank, NY 11980-9710, | rpowell-at-nanoprobes.com *
* USA | www.nanoprobes.com *
**********************************************************************






From: jim :      jim-at-proscitech.com.au
Date: Thu, 15 Jun 2000 22:57:27 +1000
Subject: RE: Epoxy cleaner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Nailon is making a very good argument for using gloves or working very
cleanly.
All epoxies I understand are "somewhat" carcinogenic. The much quoted John
Luft, years ago advised me that photographic fixer (sodium thiosulphate)
solution, chemically changed epoxies so they would not be carcinogenic. If he
was right, then first washing any body parts contaminated by epoxy resin in
photographic fixer should avert the worse. Those fixers do not dissolve or
clean epoxies.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Thursday, June 15, 2000 11:36 AM, John Nailon
[SMTP:mmjnailo-at-dingo.cc.uq.edu.au] wrote:
}
} G'day folks,
} Be very wary of any Good Epoxy Solvent, do not use it to clean epoxy off
} your skin. Anything that is a good solvent for epoxy will probably be a
} good solvent for the oils and lipids in/on your skin. These oils and
} lipids are your protection against epoxy resins entering your body.
} Remember, all epoxy resins at carcinogenic, soap and water are probably
} the safest agents to remove epoxies from your skin.
} Regards
} JVN
}
} Connie McManus wrote:
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} } At 06:09 PM 06/13/2000 -0700, Don Hammer wrote:
} } } Stuff is cheap too and if there is any left after the zillions of home
} } } uses,
} } } great on salads!!!!
} }
} } yeah, especially the used stuff........ eeeeuewwwwwwww! *G*
} }
} } connie m
} } }
} } } Don Hammer, Retired Guy
} } } ----- Original Message -----
} } } From: Connie McManus {conmac-at-cc.usu.edu}
} } } To: Tamara Howard {howard-at-cshl.org} ; Microscopy Listserver
} } } {Microscopy-at-sparc5.microscopy.com} ; Histology listserver
} } } {histonet-at-pathology.swmed.edu}
} } } Sent: Tuesday, June 13, 2000 2:29 PM
} } } Subject: Re: Epoxy cleaner?
} } }
} } }
} } } } My husband uses epoxy on a sailboat he's building. He cleans everything
} } } } up... hands, spills, etc... with plain old vinegar. We go through a LOT
} } } of
} } } } vinegar. If it's dried, then soak acetone on it until it softens, then
} } } use
} } } } vinegar (or 5% acetic acid) to mop up the residues.
} } } }
} } } } connie m
} } } }
} } } } At 10:31 AM 06/13/2000 -0400, Tamara Howard wrote:
} } } } } Back in the olden days, when BioRad sold microscopy supplies, they had
} } } } } an
} } } } } epoxy cleaner (to remove epoxies from hands, benches, etc., not made
} } } } } from
} } } } } epoxies). It was blue gunk in a little jar. Does anyone know what
} } } happened
} } } } } to this stuff? Or have an alternate?
} } } } }
} } } } } I've just been digging through catalogues and Ted Pella sells a liquid
} } } } } cleaner - any experience with it?
} } } } }
} } } } } Thanks!
} } } } }
} } } } } Tamara Howard
} } } } } CSHL
} } } } }
} } } } }
} } } } }
} } } } Connie McManus
} } } } Veterinary Diagnostics Lab
} } } } Utah State University
} } } } Logan, UT
} } } } USA
} } } }
} } } }
} } }
} } }
} } Connie McManus
} } Veterinary Diagnostics Lab
} } Utah State University
} } Logan, UT
} } USA
}
} --
} ****************************************************
} John V Nailon
} Operations Manager
} Centre for Microscopy and Microanalysis
} The University of Queensland
} St. Lucia Queensland 4072
} Phone: +61-7-3365-4214
} Fax: +61-7-3365-4422
} WWW: http://www.uq.edu.au/nanoworld/allstaff.html#Nailon
} ****************************************************





From: Smartech :      smartech-at-javanet.com
Date: Thu, 15 Jun 2000 10:02:36 -0400
Subject: SEM, How to interpret a fractured adhesive film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I needed to have a cross sectional view into an adhesive layer so that I
could count the layers, if possible. Since the adhesive acts like a highly
viscous liquid, polishing it is out of the question. Instead, I submerged
the film into liquid Nitrogen and cut it quickly w/ a scissors. The sound
of the cut was more like a breaking sound. When I examined the cross
sectional surface I observed a brain like surface. There were columns with
a highly consistent diameter and orientation. It looked crystalline. There
was zero evidence of material smearing that one would expect if one cut a
material. Is the proper interpretation that these columns existed before
the fracture and that the fracture occurred along the boundaries? Or is
the cross sectional surface generated by a rippled distortion of an highly
viscous liquid? The regularity of the surface seems to make the latter
interpretation unlikely.

I would be interested in any opinions on the interpretation of the image or
alternative means of cross sectioning the adhesive layer.

I could e-mail an image to anyone who is interested.

Thanks

Ric





From: diane.a.ciaburri-at-gdds.com
Date: 6/14/00 3:07 PM
Subject: Re: Epoxy removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Gary and others,

We haven't had a need to unpackage 'plastic' encapsulated IC's for a while but
may have a need to do so soon (preferrably not impairing functionality?!). I am
told that the last person to do this here (now retired) dripped fuming sulfuric
acid on the plastic and used frequent water rinses. We have a plasma etcher but
I was afraid it would take forever to get through the plastic.

Any hints and suggestions would be greatly appreciated.

Diane Ciaburri
Senior Materials Engineer
General Dynamics
100 Plastics Ave.
Pittsfield MA 01210

____________________Forward Header_____________________

I've used it often to unpackage epoxy microchips. These
epoxies are silicone/epoxy and not true epoxy. They are
typically called plastic packaged ICs. Whatever. I've
used MEC and DiMEC to do this. Some packages
had to be oxygen ion blasted open. Over time, the plasma
method has been much safer for the operator and the chip.
And it can make the removal process much faster. Plasma
is the current method of choice.

Whether the "epoxies" talked about here are the same as
those for IC packages is likely not to be.

gary g.






From: Hao Li :      haoli-at-glue.umd.edu
Date: Thu, 15 Jun 2000 10:40:15 -0700
Subject: information about MgAl2O4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) id JAA14804
for dist-Microscopy; Thu, 15 Jun 2000 09:46:50 -0500 (CDT)
Received: from no_more_spam.com (sparc5 [206.69.208.10])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with SMTP id JAA14801
for "MicroscopyFilteredEmail-at-msa.microscopy.com"; Thu, 15 Jun 2000 09:46:20 -0500 (CDT)
Received: from po3.glue.umd.edu (po3.glue.umd.edu [128.8.10.123])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with ESMTP id JAA14794
for {Microscopy-at-sparc5.microscopy.com} ; Thu, 15 Jun 2000 09:46:09 -0500 (CDT)
Received: from 90a280 (90a-280.umd.edu [129.2.65.24])
by po3.glue.umd.edu (8.10.1/8.10.1) with SMTP id e5FEeZE26856
for {Microscopy-at-sparc5.microscopy.com} ; Thu, 15 Jun 2000 10:40:35 -0400 (EDT)
Message-ID: {004e01bfd6f0$c3b24ac0$18410281-at-90a280}


Hi, All:

This one is not really related to microscopy. But since I am a TEM guy, I
hope I can get some help here.

Does anybody have some information about MgAl2O4 as a substrate for
perovskite films? I know it is not often used, so I am worndering if it has
a major disadvantage so that nobody is using it. Any references would be
welcome.


Thanks a lot

Hao Li





From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Thu, 15 Jun 2000 07:55:38 -0700
Subject: Polaron E5200 control board

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have a control PC board for the E5200 sputter coater.
This is the model with an Intel single chip MPU on one
end and a 4-conductor socket on the other end. The
board uses a VME connector for main interface.

Coater is trashed. Board is OK. If anybody can use
the board, first request gets it.

gary g.





From: Andy Horsewell :      horsewell-at-ipt.dtu.dk
Date: Thu, 15 Jun 2000 19:05:03 +0200
Subject: please unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please unsubscribe





From: Mary C. Pfauth :      mpfauth-at-teleport.com
Date: Thu, 15 Jun 2000 09:53:28 -0700 (PDT)
Subject: Re: autofluorescence of plant material

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Beth, more details are needed on how the tissue was processed before
viewing. I have used aniline blue to view pollen tubes in style of
Salicornia virginica which had ben fixed in Nawashin's fixative. Have also
viewed tubes of Melilotus which had simply been preserved in 70% EtOH. If
one uses glut as a fixative, it fluoresces so you won't be able to
distinguish the PT from everything else. Mary Pfauth

John P.B. & Mary
mpfauth-at-teleport.com





From: White, Woody N :      Woody.N.White-at-mcdermott.com
Date: 6/14/00 3:07 PM
Subject: Re: Epoxy removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Can't comment on sulfuric, but I have used red fuming nitric at near boiling
temperature. Apply acid, let react. Flush witn more acid, let react, etc.

Woody White

The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Gary and others,

We haven't had a need to unpackage 'plastic' encapsulated IC's for a while
but
may have a need to do so soon (preferrably not impairing functionality?!).
I am
told that the last person to do this here (now retired) dripped fuming
sulfuric
acid on the plastic and used frequent water rinses. We have a plasma etcher
but
I was afraid it would take forever to get through the plastic.

Any hints and suggestions would be greatly appreciated.

Diane Ciaburri
Senior Materials Engineer
General Dynamics
100 Plastics Ave.
Pittsfield MA 01210

____________________Forward Header_____________________

I've used it often to unpackage epoxy microchips. These
epoxies are silicone/epoxy and not true epoxy. They are
typically called plastic packaged ICs. Whatever. I've
used MEC and DiMEC to do this. Some packages
had to be oxygen ion blasted open. Over time, the plasma
method has been much safer for the operator and the chip.
And it can make the removal process much faster. Plasma
is the current method of choice.

Whether the "epoxies" talked about here are the same as
those for IC packages is likely not to be.

gary g.




From: Yan Xin :      xin-at-magnet.fsu.edu
Date: Thu, 15 Jun 2000 16:29:08 -0400
Subject: modulus for LaAlO3

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) id PAA15575
for dist-Microscopy; Thu, 15 Jun 2000 15:27:39 -0500 (CDT)
Received: from no_more_spam.com (sparc5 [206.69.208.10])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with SMTP id PAA15572
for "MicroscopyFilteredEmail-at-msa.microscopy.com"; Thu, 15 Jun 2000 15:27:08 -0500 (CDT)
Received: from magnet.fsu.edu (magnet.fsu.edu [146.201.250.2])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with ESMTP id PAA15565
for {microscopy-at-sparc5.microscopy.com} ; Thu, 15 Jun 2000 15:26:57 -0500 (CDT)
Received: from xinpc (xinpc.magnet.fsu.edu [146.201.233.133])
by magnet.fsu.edu (8.9.3+Sun/8.9.2) with ESMTP id QAA11262
for {microscopy-at-sparc5.microscopy.com} ; Thu, 15 Jun 2000 16:21:23 -0400 (EDT)
Message-Id: {4.2.0.58.20000615162741.00b5dd90-at-magnet.fsu.edu}
X-Sender: xin-at-magnet.fsu.edu
X-Mailer: QUALCOMM Windows Eudora Pro Version 4.2.0.58


Hello,

Does anyone know the shear modulus for LaAlO3?

Thanks

Yan Xin
=======================================
Yan Xin
Magnet Science & Technology
National High Magnetic Field Laboratory
Florida State University
1800 E. Paul Dirac Drive
Tallahassee, FL 32310
Tel: (850) 644 1529
Fax: (850) 644 0867
========================================







From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Thu, 15 Jun 2000 13:21:14 -0700
Subject: Re: Epoxy removal from ICs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The ion beam approach works well. I have not used it
recently on finer pitch ICs. With as-built feature sizes
of 2-4u, it is fine. It will stop at the passivation and leave
the Al bond wires intact. The resulting package looks like
it has a V-shaped pit in it (which it does). The extent of the
pit depends on the size of the die and if you want to blast
down to the lead frame or substrate.

I have not done this on finer pitch devices. I would be a
bit skeptical about these mostly because of the smaller
bond pads. The etching would still stop at the passivation.

There are numerous places in Silicon Valley that do this
on an outsource basis. Typical costs are about $75 per IC.
I can get some contacts for you if you'd like.

gary g.



At 06:55 AM 6/15/00, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Barbara Foster :      mme-at-map.com
Date: Thu, 15 Jun 2000 16:46:47 -0400
Subject: Impact of the Internet

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


MME is currently conducting research through Microscopy & Analysis
regarding the impact of the internet on microscopy and imaging facilities.
If you have not yet faxed back your responses, we'd appreciate your
participation. The questionnaire is in the center of the May issue of M&A.

Results of this survey will be reported in a Fall issue of Microscopy &
Analysis; specific information on the impact of the internet will be
presented along with data collected from other recent MME surveys and
reports from other meetings in the "Microbrew" column in the July issue of
Advanced Imaging.

Many thanks.

Best regards,
Barbara Foster
Microscopy/Marketing & Education
125 Paridon Street, Suite 102
Springfield, MA 01118-2130
PH: 413-746-6931 FX: 413-746-9311 email:mme-at-map.com

Contributing editor: Advanced Imaging "MicroBrew"





From: ERIC :      biology-at-ucla.edu
Date: Thu, 15 Jun 2000 14:29:48 -0700
Subject: Osmium Fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To the wealth of knowledge on the Microscopy list server,

I have a question about Osmium fixation..Basically I was curious to know
if there are any references to Osmium fixation at room temperature?
Everything I have seen so far only talks about fixation for 2 hours in the
refrigerator at 4 degrees...

Are there any drawback or problems that can occur if tissues specifically
Kidney and muscle would or could have? i.e. precipitation etc... etc....

Thanks in advance,

Eric
UCLA Medical Center







From: DAVID I SAXON :      DISAXON-at-prodigy.net
Date: Fri, 16 Jun 2000 03:38:26 -0700
Subject: Re: epoxy removal from ICs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


}
} We haven't had a need to unpackage 'plastic' encapsulated IC's for a while but
} may have a need to do so soon (preferrably not impairing functionality?!). I am
} told that the last person to do this here (now retired) dripped fuming sulfuric
} acid on the plastic and used frequent water rinses. We have a plasma etcher but
} I was afraid it would take forever to get through the plastic.
}
} Any hints and suggestions would be greatly appreciated.
}
} Diane Ciaburri
} Senior Materials Engineer
} General Dynamics
} 100 Plastics Ave.
} Pittsfield MA 01210

Diane,
Yes, hot fuming sulfuric and/or hot fuming nitric are used as standard
procedures for removing plastic from IC's. The process is not quite that
simple. For example, water rinses will almost certainly etch the bond pads
on the IC and thus removing connection to the outside world. Additionally,
the plastic contains fire retardants which some regions don't like being
washed down the drain. There is more detailed help through EDFAS.org (one
of ASM's branches). B&G International sells a very safe, effective etcher
which performs decapsulation automatically in minutes.

I have no association with B&G International.

David Saxon
Analytical Microscope Services
11826 Reservoir Rd. E.
Puyallup, WA 98374
253-848-7701 voice & fax
email: info-at-analyticalmicroscope.com
website: www.analyticalmicroscope.com




From: Robert Blystone :      rblyston-at-trinity.edu
Date: Thu, 15 Jun 2000 18:31:30 -0500
Subject: OnCor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To the LIST:

I am looking for information about an analysis software package by the
name of ONCOR. Does any have an adress for the company.... which may not
exist anymore?? Thanks
Blystone in Texas

Robert V. Blystone, PH.D.
Professor of Biology
Trinity University
San Antonio, Texas 78212
rblyston-at-trinity.edu
210-999-7243 FAX 210-999-7229





From: tbonner :      tbonner-at-brockport.edu
Date: Thu, 15 Jun 2000 21:14:18 -0500
Subject: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Being unable to afford a digital camera for my TEM. I'm wondering if the next
best option is to get a high end scanner to scan in negatives and then print
them on a decent printer. Any advice regarding this idea and brands of
scanners and printers that are useful? Also, what image analysis systems are
user friendly?

Dr. Thomas P. Bonner
Department of Biological Sciences
SUNY at Brockport
Brockport, NY 14420






From: Carrie Golash :      cdg126-at-psu.edu
Date: Thu, 15 Jun 2000 21:13:34 -0500
Subject: Embedding oocytes in araldite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello all -

I'm looking for advice on embedding bovine oocytes (~100 micron
diameter).
I'd like to embed them in araldite for probing with labelled lectins as
this seems to be fairly well established in the literature. The hangup is
this: we are using fairly large plastic cassettes and are having problems
losing the oocytes within the volume of araldite. Does anyone know of a way
around this? Is there something we can pre-embed the oocytes in to make a
smaller chip that we can then embed in the larger block (that is, of
course, compatible with polymerizing / clearing the araldite? Or does
anyone know of an altogether different method for oocyte embedding that is
more effective? Thanks in advance!

--Carrie Golash

Carrie Golash
John O. Almquist Research Center
Penn State University
University Park, PA 16802
W: (814) 865-5896
H: (814) 692-7926
http://www.das.psu.edu/dbrc/dbrc.htm






From: D.kierst :      dlkierst-at-config.com
Date: Fri, 16 Jun 2000 00:32:14 -0400
Subject: RE: Wanted Atomic Force Microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Our materials science microscopy lab is in need of a used atomic force
microscope. No specific model or make. All reasonable offers will be
considered. Please respond directly to Don Kierstead at or call
330-794-6600. Any help in this effort would be greatly appreciated.





From: John Twilley :      jtwilley-at-sprynet.com
Date: Fri, 16 Jun 2000 00:53:53 -0400
Subject: Re: epoxy removal from ICs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Perhaps it's time to comment on this thread.

Dichloromethane and dimethylformamide are relatively effective disrupters of most
epoxies but their action is accompanied by great swelling because the polymer becomes
engorged with the liquid before any significant solvation takes place. This will
destroy wire bonds on an IC.

Fuming (essentially anhydrous) sulfuric acid acts by the completely different process
of sulfonating reactive groups that remain on the polymer. The depolymerized and
sulfonated byproducts are quite soluble not only in the acid but usually in water as
well. The worst thing that you could do in this relatively straightforward process
is to wash with water at intervals because this would initiate almost instantaneous
corrosion. It would be advisable for a chemist, as someone trained in the handling
of reactive materials, to carry this out or at least to establish procedures and
train others with less experience. The action of sulfuric acid in this regard is
quite different than that of nitric. Nearly anhydrous nitric acid (completely
anhydrous is extremely difficult to prepare) is a very powerful oxidizer and could
lead to unstable, dangerous byproducts whereas the sulfonates resulting from the
sulfuric acid reaction are relatively stable. Water must, of course, be prevented
from splashing into any concentrated acid, especially sulfuric.

A very strong acid such as sulfuric behaves completely differently in the absence of
water. Since most acids are highly hygroscopic and are sold as water solutions, most
people do not observe this other side of their behavior. Without water to create an
ionized electrolyte, corrosion of metals will not take place. I have de-encapsulated
ICs for failure analysis in 200 degree sulfuric acid and been able to operate the IC
without replacing the .001" aluminum wirebonds that it came with. I recall one
instance where our company built prototype hybrid microelectronic circuits out of
such de-encapsulated ICs when their supplier was late getting a new design on the
market and the only ones available were already encapsulated.

The key is to realize that water must be excluded until the sulfonating acid has been
completely rinsed away by a non-aqueous liquid. As Mr. Saxon said, there are simple
and safe devices available for doing this operation. However, with proper care and
protective gear it can be done in a beaker on a hot plate in a fume hood. A few ml.s
of sulfuric acid are heated to drive off water until heavy vapors are observed over
the liquid (which may darken during heating due to trace impurities). The IC is
carefully lowered into the hot acid and a vigorous reaction ensues with the epoxy
almost instantly washing into the solution. After a few seconds the IC is then
quickly lifted out and held over a receiving vessel and flooded with a stream of
ethanol. Only after this is a final rinse in deionized water carried out, followed
by fresh electronic grade ethanol and forced drying in warm air.

The ready made devices which carry out the operation are typically a small bowl with
a hinged lid from which air is withdrawn by a gentle vacuum. An inert metal feeder
tube leads from a heated reservoir for the sulfuric acid and passes through the wall
of the bowl to a position where the encapsulated device is secured. When the lid is
closed and the slight vacuum applied, the hot acid is pulled into the bowl over the
device. It is somewhat self-limiting in that, if the lid is opened, there is no
driving force to bring more acid into the container. Naturally, the vacuum source
needs to be protected by a trap and all waste products properly handled no matter how
the procedure is carried out.

John Twilley
Conservation Scientist
(formerly, Manager of the Reliability Analysis Center,
Teledyne Microelectronics)

DAVID I SAXON wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} }
} } We haven't had a need to unpackage 'plastic' encapsulated IC's for a while but
} } may have a need to do so soon (preferrably not impairing functionality?!). I am
} } told that the last person to do this here (now retired) dripped fuming sulfuric
} } acid on the plastic and used frequent water rinses. We have a plasma etcher but
} } I was afraid it would take forever to get through the plastic.
} }
} } Any hints and suggestions would be greatly appreciated.
} }
} } Diane Ciaburri
} } Senior Materials Engineer
} } General Dynamics
} } 100 Plastics Ave.
} } Pittsfield MA 01210
}
} Diane,
} Yes, hot fuming sulfuric and/or hot fuming nitric are used as standard
} procedures for removing plastic from IC's. The process is not quite that
} simple. For example, water rinses will almost certainly etch the bond pads
} on the IC and thus removing connection to the outside world. Additionally,
} the plastic contains fire retardants which some regions don't like being
} washed down the drain. There is more detailed help through EDFAS.org (one
} of ASM's branches). B&G International sells a very safe, effective etcher
} which performs decapsulation automatically in minutes.
}
} I have no association with B&G International.
}
} David Saxon
} Analytical Microscope Services
} 11826 Reservoir Rd. E.
} Puyallup, WA 98374
} 253-848-7701 voice & fax
} email: info-at-analyticalmicroscope.com
} website: www.analyticalmicroscope.com







From: Steve Chapman :      protrain-at-emcourses.com
Date: Fri, 16 Jun 2000 10:33:15 +0100
Subject: Fractures in the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi

We often work with clients who wish to look inside materials. Firstly the
SEM is very clever it will tell you if a material is cut with a blade or a
knife or scissors!

The only way to see the true internal structure of a material is to fracture
it. Drop the material into LN2 wait until the bubbles stop and then take it
out and using heavy duty tweezers crack it.

If a material (like hair and some polymer fibres) will not crack you need to
support them in some way to make them crack. We use a water based carbon
solution and two SEM stubs. Glue the two stubs together with the water
soluble adhesive (try Spi) and then drill two or three small holes through
the stubs (about 1mm diameter). Glue the hairs together with the carbon
solution and pass then through the holes (messy). When all is dry plunge
into LN2. Tap a blade between the two stubs and ALL the material should
fracture.

Alternatively, take a fine bore drinking straw and pass the hairs plus
carbon solution into the straw. Wait until dry, dump in LN2 and flex the
straw to crack it and its contents.

Such fractures of layered materials (e.g. paints) will be best viewed in BSE
each "phase" will either be of a different contrast or fracture in a
different way. Great fun, try it?

Steve Chapman
Senior Consultant Protrain
For consultancy and professional training in EM world wide
Tel 44+ 1280 814774 Fax 814007
www.emcourses.com





From: veys-at-bota.ucl.ac.be (Pascal Veys)
Date: Fri, 16 Jun 2000 10:50:06 +0100
Subject: search for Dr. ND Hallam

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
I want to contact by Email Prof. N.D. Hallam (formerly at Melbourne and LA
Tobe - Australia)
Does anybody have his contact adress
Thanks to all in advance
Pascal

""""""""'
( O)(o )
--------------------0000--------------0000----------
Pascal VEYS
Laboratory of Plant Biology
Catholic University of Louvain
Place Croix du Sud 5 (bte 14)
B 1348 Louvain-la-Neuve
Belgium
Phone : 0032 10473004
Fax : 0032 10473471
Email : Veys-at-bota.ucl.ac.be
---------------------ooooO-----------Ooooo--------
( ) (_)(_) ( )
) ( ) (
(_) (_)






From: A.P. Alves de Matos :      apmatos-at-ip.pt
Date: Fri, 16 Jun 2000 11:44:07 +0100
Subject: Saphire Knifes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All


I«m thinking in buying a saphire knife for ultramicrotomy since they seem to
be less expensive than the traditional diamond ones. I need to cut thin
sections of sponges that are difficul to cut with glass knifes. Does anyone
have experience with these knifes? How do they compare with diamond in terms
of cutting properties and durability?


Thanks

Dr. A.P. Alves de Matos
Dental Medical School
Lisbon





From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Fri, 16 Jun 2000 08:17:39 -0500
Subject: Embedding oocytes in araldite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Many of our customers are using Agfa Duoscan scanners with excellent
results. These are "flatbed" type scanners but handle films in a separate
drawer, similar to a negative carrier in an enlarger. The advantage of this
system is not scanning through glass, eliminating the chance of Newton
Rings.
In addition, the Duoscan line offers high optical resolutions(up to
2500x2500ppi)and high dynamic ranges.
The Umax Powerlook III is also an excellent scanner where budgets may be
limited. It has 1200x2400 optical resolution and is a traditional "flatbed"
design. Also new is the Linocolor 1400 with 1200x2400 resolution with a
letter size scan bed.

Choosing a printer is more difficult, depending on your output needs. High
end photographic printers such as the Fuji Pictrography or dyesub printers
from Kodak and Sony offer top quality output but at a high price for both
hardware and cost per print. For publication quality prints, these are the
best.
Ink jet printers continue to improve in image quality, and more
importantly, long term image stability. The cost of these printers is very
low although they are very slow, and still somewhat costly per print when
used with the higher quality print materials. Most inkjets are also better
at producing color prints than monochrome prints.
Another favorite of ours is the Tektronix Phaser 850. This high quality
plain paper printer uses a unique Solid Ink technology. Ink is supplied not
in a liquid form but a solid blocks. Cost per print is very low and black
ink is free for the life of the printer.
The Phaser 850 will also handle any "office" type output such as letters
and reports with the advantage of integrating images into pages instead of
attaching all photos at the end.

George Laing
National Graphic Supply


-----Original Message-----
} From: tbonner [mailto:tbonner-at-brockport.edu]


Carrie,

We are currently working on a project involving blastocysts and since we
aren't osmicating the samples, we also have the problem of seeing them in
the resin. Embedding them in agar helps somewhat. Even though it's also
relatively transparent, the larger size of the agar chunk makes it easier to
see.

We perform our primary fixation, then buffer washes, then make a 2% agar
solution on the hot plate. When the agar cools down enough to be quite warm
to the touch (but before the gelling stage), we pipette our cells into it on
a microscope slide or cover slip, then put it into the fridge to harden. It
hardens almost immediately. Then we cut the piece of agar with the sample
into a tiny cube and continue processing it normally.

Hope this helps.

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/




-----Original Message-----
} From: Carrie Golash [mailto:cdg126-at-psu.edu]
Sent: Thursday, June 15, 2000 9:14 PM
To: Microscopy-at-sparc5.microscopy.com



Hello all -

I'm looking for advice on embedding bovine oocytes (~100 micron
diameter).
I'd like to embed them in araldite for probing with labelled lectins as
this seems to be fairly well established in the literature. The hangup is
this: we are using fairly large plastic cassettes and are having problems
losing the oocytes within the volume of araldite. Does anyone know of a way
around this? Is there something we can pre-embed the oocytes in to make a
smaller chip that we can then embed in the larger block (that is, of
course, compatible with polymerizing / clearing the araldite? Or does
anyone know of an altogether different method for oocyte embedding that is
more effective? Thanks in advance!

--Carrie Golash

Carrie Golash
John O. Almquist Research Center
Penn State University
University Park, PA 16802
W: (814) 865-5896
H: (814) 692-7926
http://www.das.psu.edu/dbrc/dbrc.htm






From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Fri, 16 Jun 2000 08:38:15 -0400
Subject: Re: Saphire Knifes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 11:44 AM +0100 6/16/0, A.P. Alves de Matos wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

*********************
I had bought a saphire knife, years ago (early 1980's). Our lab bought it
with the idea that it was a good half-way stop for a new tech who needed
something better than glass. It had certain drawbacks....the edge seemed
to collect debris and was more difficult to clean than a diamond, and of
course it wore faster too. she used it for a while (6 months?) and then
we were able to buy another diamond knife.

I haven't tried a saphire knife since, although I am partial to them in
jewelry!

Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Greg Erdos :      gwe-at-biotech.ufl.edu
Date: Fri, 16 Jun 2000 09:58:04 -0400
Subject: Re: Embedding oocytes in araldite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


When we were working with porcine oocytes it was necessary to
handle each individually so we enrobed them in agarose inside a cell of
nylon net of a dark color, using a dissecting microscope. We saved enough
of the excess nylon net to use as a "handle" to pick up the sample and
moved it from solution to solution. This should be done after fixation,
since glut fixed agarose is sometimes a problem. We also used the low temp
gelling agarose, so that we had time to work. Then put it in the frig to
solidify. It will then remain solid at room temp.
You might get better lectin labeling using one of the acrylic
resins rather than an epoxy

At 09:13 PM 06/15/2000 -0500, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Gregory W. Erdos, Ph.D.
Assistant Director, Biotechnology Program
PO Box 110580
University of Florida
Gainesville, FL 32611

Ph. 352-392-1295 Fax: 352-846-0251




From: Alan Bright :      bright-at-dial.pipex.com
Date: Fri, 16 Jun 2000 15:00:37 +0100
Subject: SEM, How to interpret a fractured adhesive film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





Dear Ric,

I see no reason why this type of cross sectional view cannot be achieved by
sectioning with a cryostat, this is the type of use our cryostats are
supplied for. If you would like more information please get back to me, I
would be happy to section some samples for you to inspect.

Best Regards

Alan Bright

Bright Instrument Co.Ltd.
St Margaret's Way
Huntingdon
PE18 6EB
England

Tel No:+44 (0)1480 454528
Fax No:+44 (0)1480 456031
Email: AlanBright-at-brightinstruments.com
Web Site: www.brightinstruments.com

++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++=

-----Original Message-----
} From: Smartech [mailto:smartech-at-javanet.com]
Sent: 15 June 2000 15:03
To: To all on the list


I needed to have a cross sectional view into an adhesive layer so that I
could count the layers, if possible. Since the adhesive acts like a highly
viscous liquid, polishing it is out of the question. Instead, I submerged
the film into liquid Nitrogen and cut it quickly w/ a scissors. The sound
of the cut was more like a breaking sound. When I examined the cross
sectional surface I observed a brain like surface. There were columns with
a highly consistent diameter and orientation. It looked crystalline. There
was zero evidence of material smearing that one would expect if one cut a
material. Is the proper interpretation that these columns existed before
the fracture and that the fracture occurred along the boundaries? Or is
the cross sectional surface generated by a rippled distortion of an highly
viscous liquid? The regularity of the surface seems to make the latter
interpretation unlikely.

I would be interested in any opinions on the interpretation of the image or
alternative means of cross sectioning the adhesive layer.

I could e-mail an image to anyone who is interested.

Thanks

Ric






From: Bruce Brinson :      brinson-at-cnst.rice.edu
Date: Fri, 16 Jun 2000 09:35:03 -0500
Subject: Re: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'll ring in & say yes. I am quite happy with this combination. You get the
digitized images with a much large field of view. Photo quality ink jets are cheap
& in general do well. You will always find extremist in on the subjects of the
infinitely best scanner & printer but here is what I bought for {9K$.
AGFA Duoscan T2500 ~$4500
500MHz PC with 1/2 Gig memory & 19" hi res monitor, CD writer ~2.5K$
Epson Stylus 870 ~$300
Photo Shop, Fovea 1.0 IP software {1K$ with student ver. of PS
Misc. supplies some $

For a MacPerson, my understanding is that in terms of image processing speed
the Macs are 2-4x faster that PC but I don't have any benchmarks on the latest
generations.

If your printing a lot, the ink jets will drain cartridges pretty quick. Down
the road I will probably pick up one of the wax printers. They cost something like
3k$ but I think it is Tektronix that offers to supply all the black wax you can
user for life, they are pretty fast & no secret papers are required (much cheap
per BW page).

Just my thoughts before coffee.

Bruce Brinson

disclaimer... no financial interest in any companies mentioned.

tbonner wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Being unable to afford a digital camera for my TEM. I'm wondering if the next
} best option is to get a high end scanner to scan in negatives and then print
} them on a decent printer. Any advice regarding this idea and brands of
} scanners and printers that are useful? Also, what image analysis systems are
} user friendly?
}
} Dr. Thomas P. Bonner
} Department of Biological Sciences
} SUNY at Brockport
} Brockport, NY 14420





From: kklos-at-mail.mse.ufl.edu
Date: Fri, 16 Jun 2000 10:36:03 -0400
Subject: Stuers Jet Polisher

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have been trying to replace the photocell on the Jetpolisher that I have
been using. It is about 15 years old and is made by Struers. The Model is
a Tenupol and the power supply is type is Polipower. Struers no longer
makes replacement parts for these units. If anyone has any information
concerning the photocells of this model (who I might contact to replace it
or the sensitivity of the photocell) it would be greatly appreciated.
Regards
Kevin





From: =shAf= :      mshaf-at-darkwing.uoregon.edu
Date: Fri, 16 Jun 2000 07:37:38 -0700
Subject: RE: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thomas writes ...

} Being unable to afford a digital camera for my TEM. I'm
} wondering if the next best option is to get a high end
} scanner to scan in negatives and then print
} them on a decent printer. ...

The next best option would be a 4x5 film scanner (~US$4k). The
problem with typical film scanners is their anticipated dynamic range
for photographic film, which is where TEM digital capture excels. I
suggest you take a representative film and evaluate the Polaroid "4x5
Ultra". Althought I'm unfamiliar with this particular 4x5 scanner, it
is the only one (I'm aware of) which is purported to scan an OD better
than 3.5 (approximately 14 f/stops ... 4 f/stops per OD unit ...
correct me if I'm wrong).
Less expensive (~US$1.2k) would be a flat bed scanner designed for
transparencies as well as hardcopy. This additional feature could be
a "drawer" for film, or a optional "lid" which provides a lamp from
above.

=shAf=





From: EBMet-at-aol.com
Date: Fri, 16 Jun 2000 11:34:47 EDT
Subject: re: Cross-section Preparation of Magnetic Recording Tape

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To All:

I need tips, descriptions, references regarding the preparation of
cross-section specimens of the magnetic layer of magnetic recording tape for
high magnfication SEM and conventional TEM observation. Thanks.

Elliot Brown
EBMET-at-AOL.COM




From: Bill Miller :      microbill-at-mohawk.net
Date: Fri, 16 Jun 2000 11:48:57 -0400
Subject: RE: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Actually the printing part has recently gotten much
easier. ElectroImage (http://www.electroimage.com) is offering new
technology that lets you print real grey scale images on simple inkjet
printers. They have grey inks and new printer drivers.

Bill Miller



At 08:47 AM 6/16/00 -0700, George Laing wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: David Henriks :      Henriks-at-CompuServe.COM
Date: Fri, 16 Jun 2000 12:11:08 -0400
Subject: re: Cross-section Preparation of Magnetic Recording Tape

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Elliot:

Off hand I know I have information on the cross sectioning of hard disk
media using the Tripod Polisher¨. I'm not sure if I have anything on
magnetic recording tape. If you send me your mailing address, I'll send
you whatever I can find that comes close.

Best regards-

David
Writing at 8:46:59 AM on 06/16/2000

***************************************************************************
************************

David Henriks
Vice President TEL: 800-728-2233
(toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499
San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com

***************************************************************************
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.

Message text written by INTERNET:"EBMet-at-aol.com"-at-sparc5.microscopy.com
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


To All:

I need tips, descriptions, references regarding the preparation of
cross-section specimens of the magnetic layer of magnetic recording tape
for
high magnfication SEM and conventional TEM observation. Thanks.

Elliot Brown
EBMET-at-AOL.COM

{





From: Edward Haller :      ehaller-at-hsc.usf.edu
Date: Fri, 16 Jun 2000 12:24:30 -0400
Subject: Osmium fix

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to Eric's question, I run a diagnostic Pathology lab at the
University of South Florida, and have a one day processing schedule for
kidney biopsies that calls for osmication in 1% buffered osmium at room
temperature for 30 minutes. I have found that for tissue pieces in the
order of 1/2 millimeter thick in one dimension the fixation is fine, and
is equivalent to our routine processing osmium fixation of one hour at 4
degrees. I haven't tried this on muscle biopsies, but if they meet the
thickness criteria they should be O.K. too,. Just make sure to rinse
these extensively (3x 10 minutes, perhaps) in buffer to remove the
excess osmium from the muscle tissue, as fluids enter and leave muscle
slower because of the extensive connective tissue sheaths around the myocytes.
Overosmication is a definite possibility, with subsequent tissue
brittleness, if tissue is left in osmium too long, no matter what the
temperature. If the tissue is too thick, uneven osmication can occur,
where the outside of the tissue is well fixed and a fixation gradient is
set up with poor fixation towards the center of the tissue. I observed
this happening at a renal lab in Pittsburgh where I used to work. Our
unstained thick sections were darker at the periphery than in the
center. Another sign of this problem is lack of specimen contrast at the
center of thin sections.
So, Eric, as far as my experience goes, it is possible to start with 4
degree, buffered osmium, and to osmicate at room temperature for a half
of an hour and get results equal to those from osmication at 4 degrees
for one hour if your tissue is sufficiently thin in at least one
dimension. I haven't noticed any precipitation problems with this
technique, nor does the osmium discolor during fixation. Our lab has
been using this technique for several years now. Take care! Ed Haller,
U.S.F. Pathology




From: David Henriks :      Henriks-at-CompuServe.COM
Date: Fri, 16 Jun 2000 12:33:25 -0400
Subject: Struers Jet Polisher

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Kevin:

You may want to ask Struers for a users list. There may be someone out
there with an older unit that is no longer being used who may be willing to
give it to you for spare parts. If they can't give you a list, let me know
- I think I can dig up an old list I put together of some previous Tenupol
users you may be able to contact. If that doesn't work, you may want to
consider upgrading to a South Bay Technology Model 550D Jet Polisher. If
you have an interest in getting more information on that option, please
contact me and I'll send you information.

Best regards-

David
Writing at 8:56:57 AM on 06/16/2000

***************************************************************************
************************

David Henriks
Vice President TEL: 800-728-2233
(toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499
San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com

***************************************************************************
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.

Message text written by
INTERNET:"kklos-at-mail.mse.ufl.edu"-at-sparc5.microscopy.com
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America



I have been trying to replace the photocell on the Jetpolisher that I have
been using. It is about 15 years old and is made by Struers. The Model is
a Tenupol and the power supply is type is Polipower. Struers no longer
makes replacement parts for these units. If anyone has any information
concerning the photocells of this model (who I might contact to replace it
or the sensitivity of the photocell) it would be greatly appreciated.
Regards
Kevin
{




From: Edward Haller :      ehaller-at-hsc.usf.edu
Date: Fri, 16 Jun 2000 12:58:34 -0400
Subject: Immunofluorescence using plastic coverslips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To the person asking about the possibility of doing immunofluorescence
microscopy on cells grown on plastic coverslips, someone has published a
technique in BioTechniques that I saved in case I needed it. Volume24,
number 6, pages 910-914, 1998 is the article titled "Mounting technique
allows observation of immuno-labeled cells on plastic coverslips".
The basic technique from M. F. Donohue et al involves using Thermanox
coverslips on which cells are grown and immunolabeled. Following
labeling, this group uses a drop of aqueous mounting medium to mount the
side of the coverslip without cells on it to a glass slide. On top of
this, the group then mounted a regular glass coverslip with an
additional drop of aqueous mountant, and then could do their microscopy.
The authors state that the inherent strong autofluorescence is greatly
reduced by this technique, and the problem with the plastic not
transmitting light well is overcome. Although I haven't tried the
technique yet, it sounds like a simple fix for a sticky problem. I hope
this is of help to you! Ed Haller, U.S.F. Pathology




From: Tom Januszewski :      tom.januszewski-at-email.swmed.edu
Date: Fri, 16 Jun 2000 11:50:10 -0500
Subject: Tissue Processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
We are currently in the market for a tissue processor for electron
microscopy. It will mainly be used for biolgical specimens (some quite
small). Although I have considerable experience with the processor sold
by RMC (Ventana), I know virtually nothing about the Lynx (now being
sold by EMS, I think). Any information on the advantages or
disadvantages of either model (or any other one that might be out there)
would be appreciated. Offline replies are welcome.

Tom Januszewski
Senior Electron Microscopist
Molecular and Cellular Imaging Facility
UT Southwestern Medical Center at Dallas
5323 Harry Hines Blvd.
Dallas, TX 75390
Email: tom.januszewski-at-email.swmed.edu





From: Glen MacDonald :      glenmac-at-u.washington.edu
Date: Fri, 16 Jun 2000 10:22:28 +0100
Subject: Re: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We are using the Agfa T2500 to scan TEM and SEM negatives, which also
gives us the ability to scan prints. The 1200 dpi is sufficient for
most needs, with 2500 dpi getting used less often and mostly with low
mag images. It is currently connected to a 233 MHz Mac G3/160 Mb RAM,
being replaced with a 400 MHz G4/320 Mb. The Agfa is driven by either a
stand alone app. or through a plug-in that runs under most software,
such as Photoshop or Object Image. You will want a lot RAM and drive
space, CDR, Ord, Jaz or DVD-RAM drives. Zip drives fill far too quickly.


Photoshop is used for publication images, although some users prefer
Canvas. Most of our image analysis uses the Object Image enhanced
version of NIH Image. It is very easy to use. I've less experience with
Image/J, but it is quickly adding capabilities and will display } 8 bits,
whereas Object Image will process 16 but only displays 8 bits. Printing
is either to an Epson 850, 3000, or our venerable Phaser IIsdx. BTW,
Tektronix sold its printer division to Xerox.

tbonner wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Being unable to afford a digital camera for my TEM. I'm wondering if the next
} best option is to get a high end scanner to scan in negatives and then print
} them on a decent printer. Any advice regarding this idea and brands of
} scanners and printers that are useful? Also, what image analysis systems are
} user friendly?
}
} Dr. Thomas P. Bonner
} Department of Biological Sciences
} SUNY at Brockport
} Brockport, NY 14420

--

Glen MacDonald
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923
glenmac-at-u.washington.edu
(206) 616-4156
(206) 616-1828 fax
***********************************************************
C:} The box said "Requires Windows95 or better". So I bought a Macintosh.
***********************************************************




From: Caroline Schooley :      schooley-at-mcn.org
Date: Fri, 16 Jun 2000 11:47:54 -0700
Subject: Re: Saphire Knifes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} I«m thinking in buying a saphire knife for ultramicrotomy since they seem to
} be less expensive than the traditional diamond ones. I need to cut thin
} sections of sponges that are difficul to cut with glass knifes. Does anyone
} have experience with these knifes? How do they compare with diamond in terms
} of cutting properties and durability?
}
} Dr. A.P. Alves de Matos
} Dental Medical School
} Lisbon

Unfortunately, you get what you pay for. Sapphire does NOT have the
durability of diamond and will be damaged by the sponge spicules. And, to
my knowldge, they can't be resharpened.


Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: Jon Ekman :      jekman-at-csd.uwm.edu
Date: Fri, 16 Jun 2000 03:18:56 -0500
Subject: Re: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


It could be a solution if you mainly want an alternative to silver prints,
or to acquire images for Powerpoint presentations. However, many
of the really useful features of a digital camera on a TEM are
unavailable if you adopt this approach, namely instant verification of
image capture, greyscale expansion, image averaging, online
analysis, and many more. Also, you still need to allocate some
space to a darkroom.

I looked at large-format transparency scanners a couple of years
ago, when there was little of this kind in the market, and concluded
that their combinations of bit depth, pixels per inch and sensitivity
and dynamic range at the high-density end of the negative (i.e.
highlight detail) was close to what was required if the objective was
merely to obtain publication quality images from a large proportion
of negative area (these images require optimised contrast and
crispness, but being small do not demand much resolution), but
really inadequate for scanning of image details (organelles,
molecules), for dense exposures and highlights, and for high
contrast subjects like replicas. Most of these scanners appeared to
be optimised for scanning positive images (large format colour
transparencies) where discrimination of detail in the extreme
shadows is not top priority. However, this becomes a major
shortcoming when dealing with negatives.

I would certainly like to know whether anyone feels that there is an
adequate solution available today.

I don't think user friendliness is the most useful criterion for
discriminating between image analysis packages. This is in any
case a fairly subjective property, depending very considerably on
the computer - literacy of the user. Most IA packages (analySIS,
Optimas, etc) are GUI-based systems, and therefore are reasonably
intuitive. One of the features that differentiates them is the balance
between the provision of off-the-peg analysis solutions and
programmability. The range of tasks demanded of an IA package is
potentially so great that there is little alternative but to evaluate them
and see if they suit your needs. However, I warn you that your
needs are likely to evolve. What seems like a simple and user-
friendly solution today will probably feel like a very limited and
inflexible one tomorrow if it has insufficient functionality and
programmability, and these things unavoidably add complexity.

Chris Jeffree

Date sent: Thu, 15 Jun 2000 21:14:18 -0500
To: Microscopy-at-sparc5.microscopy.com
} From: tbonner {tbonner-at-brockport.edu}



tbonner said:
} Being unable to afford a digital camera for my TEM. I'm wondering
} if the next best option is to get a high end scanner to scan in
} negatives and then print them on a decent printer.

Using a flat bed scanner over a digital camera for TEM images is
definately your best option budget wise. We use an Agfa Duoscan
scanner with great success. it is very versitile, we use it to scan
gels, scan TEM negatives and scan old prints. It can scan slides for
powerpoint quality presentations also. So check with Agfa to see
their latest lineup.

For printers inkjets work great when combined with photoquality
papers. Any (Epson HP & Cannon) 300dpi or higher color ink jet will
give decent images suitable for posters. For publications dye
sublimation printers work well but get one that uses a cartridge (one
piece) to replace empty media. For proofing look for laser printers
that are at least 1200dpi with extra ram (64meg on the printer is a
nice number to start with) and large toner cartridges, graphic
images burn a lot of toner. Look for a printer that will print alot
before replacement of the toner.

For software, Photoshop is a good choice. Before you spend alot on
a venders image analysis package try some of the freeware out
there like NIH image and Image J both from the NIH website. If you
can't get the free stuff to work, then spend the extra money. We use
Image J here and it works well.

For computers (Apple or PC) consider one scanner with a SCSI
interface with need a SCSI card (avoid parallel port and universal
serial bus (USB) scanners unless you like coffee breaks). So you
need one computer to hook up to the scanner but Ideally you would
also have three printers (inkjet, Laser & Dye sub or comparable) But
Inkjets are cheap make your users get one and maintain it. I bet
somewhere in your department is a networked laser printer so print
to a networked printer elsewhere. Keep the high end printer (Dye
sub/thermal printer close by) Invest in removable media that others
can use like a CD-r (HP or Plextor) and a zip drive at the bare
minimum.


Hope this helps
Jon Ekman
Associate Research Specialist
Deptartment of Biological Sciences
University of Wisconsin-Milwaukee
phone W:414.229.6471
Web1 http://www.graffitimasters.com
Web2 http://www.uwm.edu/~jekman




From: Malis, Tom :      malis-at-nrcan.gc.ca
Date: Fri, 16 Jun 2000 15:48:24 -0400
Subject: RE: Cross-section Preparation of Magnetic Recording Tape

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Embed it, Elliot and diamond knife section it in an ultramicrotome, using
the block face for the FE-SEM examination and the ultrathin sections (20-200
nm thick) for examination in the TEM. A good reference is Ho et al,
Specimen Preparation for TEM of Materials, Mat. Res. Soc. Symp. Proc., vol.
115, pp. 149-154 (MRS, Pittsburgh, 1988). You may have to search around for
someone who can do this for you, but it is far and away the best technique
for your needs.

Tom Malis
Group Leader - Characterization
Materials Technology Laboratory
Natural Resources Canada (Govt. of Canada)
568 Booth St., Ottawa, Canada
ph. 613-992-2310
FAX 613-992-8735
email: malis-at-nrcan.gc.ca


----------
From: "EBMet-at-aol.com"-at-sparc5.microscopy.com
[SMTP:"EBMet-at-aol.com"-at-sparc5.microscopy.com]
Sent: Friday, June 16, 2000 11:35 AM
To: microscopy-at-sparc5.microscopy.com
Subject: re: Cross-section Preparation of Magnetic Recording Tape


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

-----------------------------------------------------------------------.


To All:

I need tips, descriptions, references regarding the preparation of
cross-section specimens of the magnetic layer of magnetic recording
tape for
high magnfication SEM and conventional TEM observation. Thanks.

Elliot Brown
EBMET-at-AOL.COM




From: Gen Pei :      gp35-at-cornell.edu
Date: Fri, 16 Jun 2000 22:57:08 -0400
Subject: How to know the ion milling rate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


HI,

I am ion milling gold. nearly 100 nm. I want to know how long it
takes to make it thin enough to be transparent under TEM. Thanks.

Gen
******************************************************************
Gen Pei
Department of Materials Science and Engineering
Cornell University
328 Thurston Hall
tele: (607)255-5177
fax:(607)255-2365
gp35-at-cornell.edu

******************************************************************




From: cgarber-at-2spi.com
Date: Sat, 17 Jun 2000 06:52:25 -0700
Subject: RE: Re: Saphire Knifes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



For your information:

a) So far as I know they are not made any more and have not been
made for at least ten years, and

b) The economics have changed drastically from twenty years ago
when the sapphire knife did enjoy a bit of popularity. In real
terms, diamond knives, now because of the competition from Microstar
have drop significantly from what they once were, perhaps 50%,
so whatever pricing advantage there was at one time, did not
exist any more. So the Japanese company that made them discontinued
their production. It was called "Saphatome" or something like
that. Ted Pella would probably know their history, perhaps better
than I do.

Also, because of your interest in education, take a look at www.microscopy-advantage.com
. Tell me what you think. Attendees at the coming meetings
of APEM, EUREM and MSA will automatically receive a CD in their
registration materials. If you would like a copy, send me your
UPS address and I will make sure that one gets sent to you. But
it will "work" exactly as if you were on line.
--- Original Message ---
Caroline Schooley {schooley-at-mcn.org} Wrote on
Fri, 16 Jun 2000 11:47:54 -0700
------------------
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society
of America



} I´m thinking in buying a saphire knife for ultramicrotomy since
they seem to
} be less expensive than the traditional diamond ones. I need
to cut thin
} sections of sponges that are difficul to cut with glass knifes.
Does anyone
} have experience with these knifes? How do they compare with
diamond in terms
} of cutting properties and durability?
}
} Dr. A.P. Alves de Matos
} Dental Medical School
} Lisbon

Unfortunately, you get what you pay for. Sapphire does NOT have
the
durability of diamond and will be damaged by the sponge spicules.
And, to
my knowldge, they can't be resharpened.


Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html





-----
Sent using MailStart.com ( http://MailStart.Com/welcome.html )
The FREE way to access your mailbox via any web browser, anywhere!





From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Sat, 17 Jun 2000 08:11:45 -0700
Subject: Re: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 12:15 PM 6/16/00, you wrote:

} It could be a solution if you mainly want an alternative to silver prints,
} or to acquire images for Powerpoint presentations. However, many
} of the really useful features of a digital camera on a TEM are
} unavailable if you adopt this approach, namely instant verification of
} image capture, greyscale expansion, image averaging, online
} analysis, and many more. Also, you still need to allocate some
} space to a darkroom.

Why not just outsource the processing of the film? Depending
on where one resides/operates, there are typically numerous
professional and non-professional labs which will do same day
development of b/w film. I do this for 4x5 cut sheet film and
120/220 roll film from a regular camera and from the SEM recording
camera. Unless there is some overriding need or requirement for
an on-site darkroom, why not just send the film out whenever
it is needed? I could see the rationale for an on-site facility if
the TEM was producing hundreds of negs per day or perhaps
per week. Then it is a make-buy decision regarding in-house
or out-house processing.

If one is concerned about whether a shot will turn out (instant
verification), just shoot a couple more sheets or frames bracketed
around the "optimum/normal" exposure time. The cost of the
film and processing is way too low to justify a high cost digicam
for TEM. SEM imaging is of course a totally different
matter.

I find that grey scale expansion is not the sole domain of the
digicam. In a neg, additional information is there--but typically
the eye cannot see it. This is where image analysis and image
processing programs are very beneficial.


} I looked at large-format transparency scanners a couple of years
} ago, when there was little of this kind in the market, and concluded
} that their combinations of bit depth, pixels per inch and sensitivity
} and dynamic range at the high-density end of the negative (i.e.
} highlight detail) was close to what was required if the objective was
} merely to obtain publication quality images from a large proportion
} of negative area (these images require optimised contrast and
} crispness, but being small do not demand much resolution), but
} really inadequate for scanning of image details (organelles,
} molecules), for dense exposures and highlights, and for high
} contrast subjects like replicas. Most of these scanners appeared to
} be optimised for scanning positive images (large format colour
} transparencies) where discrimination of detail in the extreme
} shadows is not top priority. However, this becomes a major
} shortcoming when dealing with negatives.

Discrimination of detail in shadows is a major concern for
users of transparencies. This is why they seek high D rated
scanners. I typically scan negative and transparencies as
transmitted RGB or greyscale. This is because I find that the
scanner picks up more detail across the whole image when
scanned as a tranny.

} [snip]

} and see if they suit your needs. However, I warn you that your
} needs are likely to evolve. What seems like a simple and user-
} friendly solution today will probably feel like a very limited and
} inflexible one tomorrow if it has insufficient functionality and
} programmability, and these things unavoidably add complexity.
}
} Chris Jeffree

I agree that needs may evolve. That means that when making
the initial purchase of an image analysis program, it should
be flexible enough to allow custom augmentation. Most of
the higher end ones do. But it may turn out that one program
alone is not as good as two different programs--each being
good at different aspects of image analysis.

gary g.


~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Modern surfers use PC boards. You can too at
http://photoweb.net
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sat, 17 Jun 2000 11:50:47 -0500
Subject: Magnetic recording tape

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Tom Malis wrote:
======================================================
Embed it, Elliot and diamond knife section it in an
ultramicrotome, using
the block face for the FE-SEM examination and the ultrathin
sections (20-200
nm thick) for examination in the TEM. A good reference is Ho et
al,
Specimen Preparation for TEM of Materials, Mat. Res. Soc. Symp.
Proc., vol.
115, pp. 149-154 (MRS, Pittsburgh, 1988). You may have to search
around for
someone who can do this for you, but it is far and away the best
technique
for your needs.
======================================================

This has been our experience too, however we add the following to the
preparation protocols:

a) We coat one side of the recording tape with, say, Pt, the other side
with Al, because once in the TEM, it is important to validate that i]
nothing has fractured off during the ultramicrotomy and ii] you can keep
straight which side is which for the asymmetric cross-section. If the two
metallization lines are present in the TEM, with embedding resin on the
other side, you can be certain you are seeing the entire cross-section. If
one is missing, you might not have the entire cross-section.

b) For looking at the "faced-off-piece", we suggest an ever so slight
amount of oxygen plasma etching, in order to bring out a bit more contrast
between the ferrite or other inorganics from the matrix polymer. The
inorganics stand up like little "mesas" in the desert, giving greatly
enhanced contrast. Since you now have an element of three dimensional
nature to the same, you can gain some insight into orientation, something
that would not otherwise be possible

Disclaimer: If you are looking for someone to do this kind of work, look no
further, we have been doing this kind of sample preparation for clients on
a contract basis since the early 1970's! Our own Plasma Prep II plasma
etcher would do the described etching on the faced-off-piece after about 120
seconds of exposure.

Chuck


============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
============================================






From: Caroline Schooley :      schooley-at-mcn.org
Date: Sat, 17 Jun 2000 09:10:57 -0700
Subject: Re: Micrographs from digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

And if you're satisfied with prints at 720x1440 dpi, Epson has just made a
major leap in ink and paper longevity; read about it at
http://www.epson.com/whatsnew/ygtsi/lightfast.html
http://www.wilhelm-research.com/ . Unfortunately, the new ink cartridge
won't fit old (as in last year's) printers. Oh well - what else might I
spend $370 on?

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: Dave Audette :      deaudette-at-yahoo.com
Date: Sat, 17 Jun 2000 13:57:45 -0700 (PDT)
Subject: job posting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


List Recipients:
I am posting this message at the request of Joe Lester
and all correspondence should be sent to him.
Dave Audette
david.audette-at-sylvania.com

JOB OPENING
Scanning Electron Microscopist in Analytical
Laboratory ( Beverly, MA)

OSRAM Sylvania, Inc.
71 Cherry Hill Drive
Beverly MA 01915

DESCRIPTION:
Structural and elemental characterization of
materials used in incandescent,
fluorescent and discharge lamps, especially by
optical and electron
microscopy. Failure analyses of lamps and lighting
components. Technical
problem solving as a member of a team. Oral and
written communication of
results and conclusions with client population.

POSITION REQUIREMENTS:
Competence in optical and electron microscopy of
materials including EDS. An
understanding of failure analysis. Ability to work
independently and/or in a
team and to communicate effectively.

EDUCATION AND EXPERIENCE REQUIREMENTS:
B.S., or higher, in Materials Science,
Chemistry, or Physics. 2-5
years experience in SEM/EDS. Experience with lamp
components is desirable.


Please send a resume to


Dr. Joe Lester
Technical Assistance Lab
OSRAM Sylvania Inc.
71 Cherry Hill Drive
Beverly, MA 01915
e-mail: joe.lester-at-sylvania.com



__________________________________________________
Do You Yahoo!?
Send instant messages with Yahoo! Messenger.
http://im.yahoo.com/




From: Nestor J. Zaluzec :      zaluzec-at-sparc5.microscopy.com
Date: Sun, 18 Jun 2000 19:04:44 -0500
Subject: Administrivia: Listserver Archives Updated.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues....

I had fallen behind on the Archive Updates. They are now
current through May 31, 2000.

http://www.msa.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Nestor
Your Friendly Neighborhood SysOp.







From: John Nailon :      mmjnailo-at-dingo.cc.uq.edu.au
Date: Mon, 19 Jun 2000 10:23:25 +1000
Subject: Re: Saphire Knifes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


G'day All,
In my experience Sapphire Knives are an excellent replacement for glass
knives when working with soft materials. Sapphires are much softer than
diamond and are more easily damaged than diamond. Sponges are NOT soft
tissue, they contain very hard inorganic salt spicules that damage both
glass and sapphire knives.
Regards
JVN

Leona Cohen-Gould wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} At 11:44 AM +0100 6/16/0, A.P. Alves de Matos wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Dear All
} }
} }
} } I«m thinking in buying a saphire knife for ultramicrotomy since they seem to
} } be less expensive than the traditional diamond ones. I need to cut thin
} } sections of sponges that are difficul to cut with glass knifes. Does anyone
} } have experience with these knifes? How do they compare with diamond in terms
} } of cutting properties and durability?
} }
} }
} } Thanks
} }
} } Dr. A.P. Alves de Matos
} } Dental Medical School
} } Lisbon
}
} *********************
} I had bought a saphire knife, years ago (early 1980's). Our lab bought it
} with the idea that it was a good half-way stop for a new tech who needed
} something better than glass. It had certain drawbacks....the edge seemed
} to collect debris and was more difficult to clean than a diamond, and of
} course it wore faster too. she used it for a while (6 months?) and then
} we were able to buy another diamond knife.
}
} I haven't tried a saphire knife since, although I am partial to them in
} jewelry!
}
} Lee
}
} Leona Cohen-Gould, M.S.
} Sr. Staff Associate
} Director, Electron Microscopy Core Facility
} Manager, Optical Microscopy Core Facility
} Joan & Sanford I. Weill Medical College
} of Cornell University
} voice (212)746-6146
} fax (212)746-8175

--
****************************************************
John V Nailon
Operations Manager
Centre for Microscopy and Microanalysis
The University of Queensland
St. Lucia Queensland 4072
Phone: +61-7-3365-4214
Fax: +61-7-3365-4422
WWW: http://www.uq.edu.au/nanoworld/allstaff.html#Nailon
****************************************************




From: XUELI ZHAO :      xzhao-at-unlserve.unl.edu
Date: Sun, 18 Jun 2000 21:10:49 -0500 (CDT)
Subject: Sight seeing while EM meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, Every one:

I had a argument with my husband, he says he is going to the
Kunming for an EM meeting,

http://www.iphy.ac.cn/microsc/IKSM.html

He says there are many good scientists going which I believe with doubt.
But I think he is going for a sight seeing. There is a Chinese saying:
The mountains and waters in Guilin are the most beautiful ones under the
sky. I have asked him to bring me, he agreed but unable to get the same
airline ticket (UA fully booked). Any body is going and knows alternative
airlines, please contact me. Thanks a lot.

Xueli






From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Sun, 18 Jun 2000 20:55:08 -0700
Subject: Warning about DVD-RAM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have two DVD-RAM drives and 15 media (Type I and II).
I have found that these are riddled with write & read errors.
Be careful when using this storage media.

My main unit is a Panasonic LF-D101 (SCSI) and
the second one is the same. The third is a Matshushita
ID unit.

Scandisk will report either many errors that are fixed or
no errors. Either way, the media/drive will write faulty
file contents.

Be careful.






From: ComCryLab1-at-aol.com
Date: Mon, 19 Jun 2000 10:27:57 EDT
Subject: CRYSTAL: MgAl2O4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Hao,
We have in stock (100), (110) and (111) MgAl2O4 substrates with 7 angstrom
finish ready for laser ablation or other kinds of epitaxial film deposition.
Contact me for info regarding perovskite lattice matching.
Best regards,
Mike Urbanik
www.crystalguru.com

{ { Subj: information about MgAl2O4
Date: 6/15/00 3:33:44 PM Eastern Daylight Time
From: haoli-at-glue.umd.edu (Hao Li)
To: Microscopy-at-sparc5.microscopy.com

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America
To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
-----------------------------------------------------------------------.


Hi, All:

This one is not really related to microscopy. But since I am a TEM guy, I
hope I can get some help here.

Does anybody have some information about MgAl2O4 as a substrate for
perovskite films? I know it is not often used, so I am worndering if it has
a major disadvantage so that nobody is using it. Any references would be
welcome.


Thanks a lot

Hao Li

} }




From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Mon, 19 Jun 2000 09:46:49 -0400
Subject: TEM of yeast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,
I processed a series of pellets of yeast for a client using a glut-pfa fix,
osmium, dehydration through ethanols, and a day & a half step-wise
infiltration into Spurr's resin (1:1 with Ethanol for 4 hr, pure resin
overnight under light vacuum, then fresh for 3 more hours, then embed in
fresh). Polymerization was overnight at 60C. About half the blocks were
soft and had to be returned to the oven for a prolonged polym. (over the
weekend). I was able to get sections from each of the 10 samples, but in
the 'scope, many of these looked a bit like swiss cheese. those that were
not lacy exhibited areas where the resin pulled away from the cell coats of
the yeasts. Clearly something went wrong with the
infiltration/polymerization. I used the same batch of resin for other
things, and its fine.

Does anyone out there have experience with yeast? Any suggestions?
I'd like to give this peson some usable data!

Thanks in advance,
Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Mon, 19 Jun 2000 10:09:46 -0500
Subject: TEM of yeast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Lee,

My only thought is that your infiltration/dehydration were too short, or
your 100% ethanol had absorbed too much water. My understanding is that
Spurr's is very sensitive to small quantities of water, with soft blocks and
holes being symptoms of incomplete dehydration. Sometimes we extend the
dehydration through 3 changes of 100% ETOH with molecular sieves, then on
through 2-3 changes of propylene oxide. Infiltration is usually 1:2
PO:Resin, followed by 1:1, 2:1, then a couple changes of pure resin
overnight or for 4-8 hours, then final embedding in another change of pure
resin.

I don't know if yeast is more problematic than other things in this respect,
however.

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/




-----Original Message-----
} From: Leona Cohen-Gould [mailto:lcgould-at-mail.med.cornell.edu]
Sent: Monday, June 19, 2000 8:47 AM
To: Microscopy-at-sparc5.microscopy.com


Hi All,
I processed a series of pellets of yeast for a client using a glut-pfa fix,
osmium, dehydration through ethanols, and a day & a half step-wise
infiltration into Spurr's resin (1:1 with Ethanol for 4 hr, pure resin
overnight under light vacuum, then fresh for 3 more hours, then embed in
fresh). Polymerization was overnight at 60C. About half the blocks were
soft and had to be returned to the oven for a prolonged polym. (over the
weekend). I was able to get sections from each of the 10 samples, but in
the 'scope, many of these looked a bit like swiss cheese. those that were
not lacy exhibited areas where the resin pulled away from the cell coats of
the yeasts. Clearly something went wrong with the
infiltration/polymerization. I used the same batch of resin for other
things, and its fine.

Does anyone out there have experience with yeast? Any suggestions?
I'd like to give this peson some usable data!

Thanks in advance,
Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Mon, 19 Jun 2000 09:17:18 -0700
Subject: Re: How to know the ion milling rate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Gen Pei,
I have looked at gold contacts, lifted from an electronic device, that were
supposed to be 100 nm thick. I could see the structure clearly at 200 kV.
At 10:57 PM 6/16/00 -0400, you wrote:

} HI,
}
} I am ion milling gold. nearly 100 nm. I want to know how long it
} takes to make it thin enough to be transparent under TEM. Thanks.
}
} Gen

Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca





From: jmkrupp-at-cats.ucsc.edu (Jon Krupp)
Date: Mon, 19 Jun 2000 11:25:24 -0700
Subject: Formvar & 1,2 dichloroethane

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi:

I was just visited by one of our EH&S folks who wanted to know why I had
1,2 dichloroethane.

Seems they track purchases and I bought some last year to make formvar films.

1,2 dichloroethane is on their bad list as a carcinogen. He was actually
here to figure out how much might be going up the fume hood so he could
make a report to the local air quality agency. But as we talked, it seemed
like it would be better to not have the stuff around.

Anyone have experience with formvar in chloroform? I read it works but have
never tried it. According to our EH&S guys, chloroform would be better than
1,2, dichloroethane.

BTW we are in California and must abide by some pretty strict rules, it may
seem like they are going overboard, but they are just trying to do their
job.

Jonathan Krupp
Microscopy & Imaging Lab
University of California
Santa Cruz, CA 95064
(831) 459-2477
jmkrupp-at-cats.ucsc.edu






From: Connie A Cummings :      rosscac-at-okstate.edu
Date: Mon, 19 Jun 2000 12:42:17 -0500
Subject: Sperm Processing for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


{P} Good day to all on the listserver, {/P} {P}   {/P} {P} I have a person in my department who is interested in processing horse sperm for TEM.  Does anyone who does this routinely be willing to give me some pointers as to how to process them?  I've only worked with muscle and brain tissue so this is kinda new - I have processed cells from cell culture for TEM would it be the same procedure? {/P} {P}   {/P} {P} Thanks so much, {/P} {P} Connie Cummings, DVM {/P} {P} Instructor Anatomic Pathology {/P} {P} Department VBP {/P} {P} Oklahoma State University {/P} {P} {/P}

{/html}



From: Mike Coviello :      coviello-at-mae.uta.edu
Date: Mon, 19 Jun 2000 15:12:30 -0500
Subject: TEM-looking for a for-hire independent FIB company that does TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




---------- Forwarded message ----------


Hi Y'all:
We are looking for a for-hire independent FIB company that has
experience in preparing TEM cross-sections of semiconductors. Please
contact me if you do this, or know of a lab that does.
Regards,
Michael Coviello
Lab Manager
Materials Science
University of Texas at Arlington





From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Mon, 19 Jun 2000 16:06:34 -0400
Subject: RE: TEM of yeast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to everyone for your helpful comments. I'm taking another stab at
it with smaller pellets, longer times and the addition of prop. ox. steps
after the ethanol.
I had pretty much decided to do all that anyway, but its nice to gets
confirmation ofone's ideas!

Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Jean-Pierre Slakmon :      slakmon-at-soquelec.com
Date: Mon, 19 Jun 2000 17:56:55 -0400
Subject: TEM-looking for a for-hire independent FIB company that does

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Try FIBICS in Ottawa, Canada. Contacts are Mike Phaneuf or Louise Weaver.
Their telephone number is 613-860-0861.
email: mphaneuf-at-fibics.com or lweaver-at-fibics.com

Jean-Pierre Slakmon
Soquelec Limited
5757 Cavendish Boulevard, Suite 101
Montreal, Quebec
Canada H4W 2W8
Tel: 514-482-6427 / Fax: 514-482-1929
URL: http://www.soquelec.com



-----Original Message-----
} From: Mike Coviello [mailto:coviello-at-mae.uta.edu]
Sent: Monday, June 19, 2000 4:12 PM
To: listserver


Hi Y'all:
We are looking for a for-hire independent FIB company that has
experience in preparing TEM cross-sections of semiconductors. Please
contact me if you do this, or know of a lab that does.
Regards,
Michael Coviello
Lab Manager
Materials Science
University of Texas at Arlington






From: Frank Macaluso :      macaluso-at-aecom.yu.edu
Date: Mon, 19 Jun 2000 18:00:27 -0400
Subject: Re: TEM of yeast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Lee,
We follow a very similar protocol to yours with the exception that we use
propylene oxide rather than ethanol for infiltration.
PO : Spurrs 1:1 1 hour
PO : Spurrs 1:3 1- 2 hours
Spurrs 1 - 2 hours
Spurrs overnight (we do not infiltrate under vacuum)
fresh Spurrs 1 - 2 hours
polymerization at 60C 48 hours

We had the problem you describe when we tried to embed yeast in EPON
equivalents. Switching to Spurrs was the fix for us.
Frank

At 09:46 AM 6/19/00 -0400, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Jon Mulholland :      jwm-at-genome.stanford.edu
Date: Mon, 19 Jun 2000 16:23:07 -0700 (PDT)
Subject: Re: TEM of yeast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Leona,
The standard way for dealing with yeast cells is to remove the
cell wall. This is done with glusulase (fancy name for snail guts) and or
lyticase (both from Sigma). If the wall cannot be removed for
experimental reasons (i.e. study of the plasma membrane/cell wall
interface) then you need to modify the carbohydrate linkages of the cell
wall to make the wall more permeable. This can be done by treating the
cells, after fixation, with 1% sodium metaperiodate for about 15 minutes.
However, most researchers simply wishing to examine yeast morphology
remove the wall because this not only improves inflitration of the resin
it also allows more extraction of the cytoplasm and thus makes it easier
to resolve structures and membranes within the ribosome rich yeast cell. A
classic protocol, by Byers and Goetsch, can be found in Vol 194, Methods
in Enzymology (AKA Guthrie and Fink) pg 602. I strongly encourage you to
read this article as yeast can be very problematic. You also want to use
the Hard Spurrs formulation and use 100% acetone (or propylene oxide) as
the last dehydration step before going into 1:1 resin. I have not
observed any difference in the ultrastructure of cells embedded in Spurrs
vs. polybed 812 (when the wall is removed).
If you want to do immuno-EM you might find our protocol useful;
checkout:
http://genome-www.stanford.edu/group/botlab/protocols/EM_protocol.pdf.

Jon Mulholland
Genetics Dept
Stanford University School of Medicine
Stanford, CA 94305-5120


On Mon, 19 Jun 2000, Leona Cohen-Gould wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi All,
} I processed a series of pellets of yeast for a client using a glut-pfa fix,
} osmium, dehydration through ethanols, and a day & a half step-wise
} infiltration into Spurr's resin (1:1 with Ethanol for 4 hr, pure resin
} overnight under light vacuum, then fresh for 3 more hours, then embed in
} fresh). Polymerization was overnight at 60C. About half the blocks were
} soft and had to be returned to the oven for a prolonged polym. (over the
} weekend). I was able to get sections from each of the 10 samples, but in
} the 'scope, many of these looked a bit like swiss cheese. those that were
} not lacy exhibited areas where the resin pulled away from the cell coats of
} the yeasts. Clearly something went wrong with the
} infiltration/polymerization. I used the same batch of resin for other
} things, and its fine.
}
} Does anyone out there have experience with yeast? Any suggestions?
} I'd like to give this peson some usable data!
}
} Thanks in advance,
} Lee
}
} Leona Cohen-Gould, M.S.
} Sr. Staff Associate
} Director, Electron Microscopy Core Facility
} Manager, Optical Microscopy Core Facility
} Joan & Sanford I. Weill Medical College
} of Cornell University
} voice (212)746-6146
} fax (212)746-8175
}
}
}





From: Keith Ryan :      kpr-at-ccms.ac.uk
Date: Tue, 20 Jun 2000 08:16:55 +0100
Subject: Re: Formvar & 1,2 dichloroethane

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jonathan

Formvar in chloroform works well. I don't think we have ever tried
dochloroethane.

I have had to do the job myself recently and made a minor discovery -
the film seems to stick very well to acid washed slides! So, not so
cleverly clean. Then the second trick - float the film soon after it
has dried, within a minute. Then it seems to work better.

Keith

_______________________
Keith Ryan (Dr)
Marine Biological Association
Citadel Hill
Plymouth
Devon PL1 2PB
England

Tel. 0044 (0)1752 633279 (with answer machine)
also 0044 (0)1752 633249
Fax. 0044 (0)1752 633102

e-mail: kpr-at-wpo.nerc.ac.uk




From: Ian MacLaren :      maclariz-at-yahoo.co.uk
Date: Tue, 20 Jun 2000 08:40:52 +0100 (BST)
Subject: RE: Ceramic cross-section preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,
Thanks to all who responded to my enquiry and gave me useful advice concerning
the preparation of cross-sections of ceramic thin films.

Hopefully, I should now be able to prepare some better thin film specimens
using one or more of the suggestions that I received.

Thanks again

Best wishes

=====
Ian MacLaren
Beijing Laboratory of Electron Microscopy
Chinese Academy of Sciences, P.O. Box 2724
100080 Beijing
China
General Email: ian.maclaren-at-physics.org
Work (esp. large attachments): maclaren-at-image.blem.ac.cn

____________________________________________________________
Do You Yahoo!?
Get your free -at-yahoo.co.uk address at http://mail.yahoo.co.uk
or your free -at-yahoo.ie address at http://mail.yahoo.ie




From: Krzysztof Jan Huebner :      hubner-at-IOd.krakow.pl
Date: Tue, 20 Jun 2000 11:34:53 +0200 (MET DST)
Subject: SCIA2001, First Announcement and Call for Papers: (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



12th Scandinavian Conference
on Image Analysis
SCIA 2001

June 11-14, 2001 in Bergen, Norway

Sponsored by: IAPR (The International Association
for Pattern Recognition)
http://www.iapr.org

First Announcement and Call for Papers:

http://www.ux.his.no/scia2001/

Invitation to the 12th SCIA.
Following the previous conferences in Greenland, SCIA 2001 -
the 12th Scandinavian Conference on Image Analysis - will be
held in Bergen on the west coast of Norway. The conference is
arranged by the Norwegian Society for Image Processing and
Pattern Recognition (NOBIM) and sponsored by the International
Association for Pattern Recognition (IAPR). The conference
venue is Grieghallen, located in the city centre.

Scientific Program:
The conference will offer internationally acclaimed speakers in
plenary talks and parallel sessions with selected oral presentations
and posters. The conference language is English.
The different presentations will cover unpublished theoretical or
applied research results.

Invited Speakers:
Professor Theo Pavilidis, State University of New York at Stony
Brook, USA: "History of Image Analysis"
Professor Josef Begün, University of Halmstad, Sweden:
"Biometric Person Authentication"
Professor Matti Pietikäinen, University of Oulu, Finland:
"Machine Vision and Media Processing"
Associate Professor Torbjřrn Eltoft, University of Tromsř, Norway:
"Neural Network approaches to Cluster-Detection-and-Labelling"


In addition, we are working to find an invited speaker for the
subject: "Images in the future mobile terminals".

Pre-conference Workshop/Tutorial:
A set of pre-conference half-day workshops/tutorials will be held
on June 11, 2001:
1. ICA (Independent Component Analysis): Professor Erkki Oja,
Helsinki University of Technology,
Finland.
2. Data fusion: Professor Jon Atli Benediktsson,
University of Iceland.

Conference topics:
* Image analysis
* Computer vision
* Pattern recognition
* Neural networks
* Statistical methods
* Industrial applications
* Multimedia
* Biomedical applications
* Remote sensing
* Future technologies

Paper submission and registration for presenting authors:
Only full papers in English will be accepted, and the length should
not exceed eight pages. All papers will be refereed by two
reviewers for publication in the conference proceedings.
Please send four copies of your paper to:

SCIA2001,
Department of Electrical and Computer Engineering,
Stavanger University College,
P.O.Box 2557 Ullandhaug,
N-4091 Stavanger,
Norway

Important dates:

Paper submission deadline: November 6, 2000
Notification of acceptance: January 19, 2001
*Camera-ready copy: March 19, 2001

*Camera-ready copy must be accompanied by registration
and payment by presenting author.
The cover page must contain:
* Title of the paper
* Name(s), complete address and e-mail for the author(s)
* Brief abstract (150-200 words)
* Keywords describing the main subject of the paper (3-5 words)

* Author's opinion on whether the paper is most suitable for oral
or poster presentation
* Name and address for correspondence

Papers considerably longer than the final size, risk being rejected.
The fee for one or two extra pages is NOK 500 per page.
The fee for colour illustrations is NOK 4000 per page.
The decision on oral or poster presentation will be taken solely on
suitability, not on paper quality. Paper submission information is
available at http://www.ux.his.no/scia2001/, where the LaTeX style
file and an example file in the recommended two-column LaTeX
format is available.

Enquires:
If you have scientific questions, please contact
Ivar.Austvoll-at-tn.his.no. The program with further information about
registration and payment will be send to you February 1, 2001.
If you occasionally have seen this announcement, and want to have
the program sent to you, please contact scia-at-plus-convention.no.

Program Committee:
Dr. Ivar Austvoll (Chairman), Norway
Prof. Jussi Parkkinen, Finland
Prof. Fritz Albregtsen, Norway
Dr. Alfred Hanssen, Norway
Prof. Gunilla Borgefors, Sweden
Dr. Anne Solberg, Norway
Dr. Bjarne Ersbřll, Danmark

Organized by: Norsk forening for bildebehandling og
mřnstergjenkjenning (NOBIM)
(Norwegian Society for Image Processing and
Pattern Recognition) http://www.nobim.no/

SCIA2001 web site:

http://www.his.no/scia2001/ or http://www.ux.his.no/scia2001/







From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Tue, 20 Jun 2000 08:54:03 -0400
Subject: Re: Sperm Processing for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 12:42 PM -0500 6/19/0, Connie A Cummings wrote:

} Good day to all on the listserver,
}
} I have a person in my department who is interested in processing horse
} sperm for TEM. Does anyone who does this routinely be willing to give me
} some pointers as to how to process them? I've only worked with muscle and
} brain tissue so this is kinda new - I have processed cells from cell
} culture for TEM would it be the same procedure?
}
} Thanks so much,
} Connie Cummings, DVM
} Instructor Anatomic Pathology
} Department VBP
} Oklahoma State University
****************************
Connie,
Assuming that your colleague will bring you a semen sample, and that
orientation is not critical, you can spin the sperm to a pellet and treat
it like any other cell pellet. I've done various rodent, marsupial and
human sperm samples this way. Once at the microscope, you will have to
hunt around a bit to fine the appropriate views (head, mid-piece,
tails,etc), but since there are so many cells in the pellet, I've always
found what we wre looking for. Its the easiest way to go.
Good luck,
Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Paula Allan-Wojtas :      AllanWojtasP-at-Em.agr.ca
Date: Tue, 20 Jun 2000 10:09:25 -0400
Subject: power problems with the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, all,

I am posting this question to see if we can get some help for our building engineer to deal with problems we are having with our SEM (JEOL T330A).

We suspect that we are having a problem with the building power supply to our SEM. Other folks in our building using other types of equipment have found it necessary to use power conditioners or UPS systems to run their equipment - these systems seem to do the trick for them.

We are trying this approach with the power supply to our SEM, but find that there are some problems. When using the UPS system, we have to connect it up, put it on bypass, start up the SEM, then switch the UPS over after that - a real nuisance. If we don't do this, the SEM won't start.

Does anyone else have this problem, and if so, has found an easy solution? Please contact me offline if you have any suggestions and other words of wisdom. Thanks in advance. Also, thanks for all the help you as a group have given me when I posted previous questions - sometimes I forget to say thank you.

Paula.

Paula Allan-Wojtas
Research Scientist - Food Microstructure
Agriculture and Agri-Food Canada
Atlantic Food and Horticulture Research Centre
Kentville, Nova Scotia Canada B4N 1J5

Tel: (902) 679-5566
FAX: (902) 679-2311

email: allanwojtasp-at-em.agr.ca





From: John F. Mansfield :      jfmjfm-at-engin.umich.edu
Date: Tue, 20 Jun 2000 10:13:28 -0400
Subject: A Gatan User Meeting at M&M Philadelphia

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



This is a preliminary message to all Gatan users who will be
attending the Microscopy and Microanalysis Meeting in Philadelphia
this August (13th to the 17th). I think it has become time to form a
User's Group to discuss the level of service and support that we are
receiving from Gatan. We should determine where the company should
be focusing its efforts and lobby them to correct problems that are
most important to us. Please let me know if you are interested in
attending such a meeting so I can gauge whether it should be held,
and how large a conference room I would need to reserve in Philly.

Note, Gatan representatives are encouraged to attend this meeting,
but it will be a user meeting run by the users.

I will not rant and rave here in a completely open forum, as I
believe it would be unfair. If you have concerns or comments on this
subject, whether or not you are attending M&M2000, please contact me
directly. Do NOT reply to the list, check the "To:" header before
sending your message, it should say "jfmjfm-at-engin.umich.edu" only.

Thank You.

John Mansfield.

Disclaimer: Opinions expressed in this message are my own personal
ones and do not represent necessarily those of my employers.

--

Dr. John Mansfield CPhys MInstP
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (734) 936-3352 FAX (734) 763-2282
Cellular Phone: (734) 358-7555
(Leaving a phone message at 936-3352 is preferable to 358-7555)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html
Location: Lat. 42ˇ 16' 48" Long. 83ˇ 43' 48"




From: David H. Hall :      hall-at-aecom.yu.edu
Date: Tue, 20 Jun 2000 11:52:26 -0400
Subject: acetone + LR Gold

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have experience taking specimens from acetone into LR Gold
resin? Some of the brochures on LR White seem to recommend against it, but
we are hoping it may be OK for LR Gold. We are doing freeze substitution
through acetone for EM-immunocytochemistry.

Thanks in advance for your help.
David H. Hall
Center for C. elegans Anatomy
Department of Neuroscience
1410 Pelham Parkway
Albert Einstein College of Medicine
Bronx, NY 10461

phone (718) 430-2195 FAX (718) 430-8821
hall-at-aecom.yu.edu
website: www.aecom.yu.edu/wormem




From: White, Woody N :      Woody.N.White-at-mcdermott.com
Date: Tue, 20 Jun 2000 13:17:00 -0500
Subject: Re:power problems with the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Paula,

Sounds like the UPS capacity (VA) is sufficient to run the system, but under
rated for the start-up surge. One solution would be a higher capacity UPS.
One
sized to handle the surge load, however, could be quite large and expensive.

If the nature of your problems is related to line noise, but not voltage
levels,
you might consider an "ultra isolation transformer" and experiment with
various
grounding options to minimize interference.

If voltage flucuations are the problem, investigate a "ferro resonant"
transformer to stabilize the line. This transformer *should* be somewhat
more
economical than a similar capacity UPS. BEWARE: These devices produce a
loud
hum. You don't want it in the same room without some sort of noise
attenuation.


.Haven't checked relative pricing, but here is a typical link:
http://www.sola-hevi-duty.com/

Another posibility... Can you seperate the vacuum pump(s) supply from the
electronics, powering the pumps directly and using your UPS for only the
electronics? That may lessen the start-up load enough for the UPS to
funcion
normally.

Woody White
McDermott Technology, Inc.

The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hi, all,

I am posting this question to see if we can get some help for our building
engineer to deal with problems we are having with our SEM (JEOL T330A).

We suspect that we are having a problem with the building power supply to
our
SEM. Other folks in our building using other types of equipment have found
it
necessary to use power conditioners or UPS systems to run their equipment -
these systems seem to do the trick for them.

We are trying this approach with the power supply to our SEM, but find that
there are some problems. When using the UPS system, we have to connect it
up,
put it on bypass, start up the SEM, then switch the UPS over after that - a
real
nuisance. If we don't do this, the SEM won't start.

Does anyone else have this problem, and if so, has found an easy solution?
Please contact me offline if you have any suggestions and other words of
wisdom.
Thanks in advance. Also, thanks for all the help you as a group have given
me
when I posted previous questions - sometimes I forget to say thank you.

Paula.

Paula Allan-Wojtas
Research Scientist - Food Microstructure
Agriculture and Agri-Food Canada
Atlantic Food and Horticulture Research Centre
Kentville, Nova Scotia Canada B4N 1J5

Tel: (902) 679-5566
FAX: (902) 679-2311

email: allanwojtasp-at-em.agr.ca




From: tellis2-at-hallmark.com
Date: Tue, 20 Jun 2000 13:56:33 -0500
Subject: EDS Window

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


HI:
For the second year in a row I am having to have our thin window (low
element type-brand name with held) replaced.
We have the EDS connected to a TOPCON SM-500 environmental SEM. I look at a
lot industrial dusts for particle size and composition and the manuf. of
the EDS system says they poke holes in the polymer window.
Has anyone else have this problem?
I have elected to have a thin beryllium installed this time since my
boss is upset about spending $7000 every year plus the downtime and since
low element detection is not critical .
Thanks
Terry Ellis
Hallmark Cards Inc.





From: carla_aiwohi-at-usgs.gov
Date: Tue, 20 Jun 2000 13:37:19 -0700
Subject: TEM cell pellets in agarose

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,

Does anyone have suggestions for encapsulating cell pellets in agarose?

I am working with flatfish cells but the pellets don't look cohesive enough
to withstand washing, dehydration etc.
We have some Type I agarose (Sigma, gel temp 36C, melting temp. 86C). I
plan to post-fix in 1% osmium tetroxide followed by 1% aqueous uranyl
acetate then embed in Spurr's resin.

As this is my first time working with cells, any help will be greatly
appreciated.

Thank you very much,
Carla Aiwohi
Western Fisheries Research Center
Seattle, WA





From: White, Woody N :      Woody.N.White-at-mcdermott.com
Date: Tue, 20 Jun 2000 16:57:00 -0500
Subject: Re:EDS Window

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Terry,

The particulate (have also heard) can "shoot" holes in a thin window. My
former
SEM/EDS was an Etec with a LARGE roughing system and a turret detector with
Be,
UTW, and "open" positions. Both windows lasted the 15 odd years it was in
use
before replacement. An important point is to evacuate and vent slowly so as
to
not accelerate the particles into the window.

I added a manual valve to the Etec between the roughing system and the
chamber.
With the automatic valves closed, I could slowly open the manual valve to
rough
the chamber down to a point where any particulate is not disturbed, close
the
valve, then switch to automatic to finish the evacuation. Venting gas was
from
my cryo of liquid nitrogen (pretty dry!) which I pressure/flow controlled
for
similar results on venting.

My modification was driven by the need to avoid damage to fluffy ceramic
specimens, but worked well to protect the detector also.

Have not added such a feature to the new SEM/Thin Window EDS system, but
(for
now) the chamber is pristine and the fluffy specimens are not part of the
work
mix.

Woody White
McDermott Technology, Inc

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


HI:
For the second year in a row I am having to have our thin window (low
element type-brand name with held) replaced.
We have the EDS connected to a TOPCON SM-500 environmental SEM. I look at a
lot industrial dusts for particle size and composition and the manuf. of
the EDS system says they poke holes in the polymer window.
Has anyone else have this problem?
I have elected to have a thin beryllium installed this time since my
boss is upset about spending $7000 every year plus the downtime and since
low element detection is not critical .
Thanks
Terry Ellis
Hallmark Cards Inc.




From: Earl Weltmer :      earlw-at-pacbell.net
Date: Tue, 20 Jun 2000 15:50:36 -0700
Subject: Re: power problems with the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

My guess is that the SEM inititally draws power that exceeds the ratings of the UPS.
It is probably due to the rotary pumps that can take up to ten amps upon start up.

The alternatives are to increase the power rating of the UPS or rewire the rotary pump so it draws it's power from the line and not through the UPS (the pumps is not that sensitive anyway).

Good Luck,

Earl weltmer

Paula Allan-Wojtas wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hi, all,
}
} I am posting this question to see if we can get some help for our building engineer to deal with problems we are having with our SEM (JEOL T330A).
}
} We suspect that we are having a problem with the building power supply to our SEM. Other folks in our building using other types of equipment have found it necessary to use power conditioners or UPS systems to run their equipment - these systems seem to do the trick for them.
}
} We are trying this approach with the power supply to our SEM, but find that there are some problems. When using the UPS system, we have to connect it up, put it on bypass, start up the SEM, then switch the UPS over after that - a real nuisance. If we don't do this, the SEM won't start.
}
} Does anyone else have this problem, and if so, has found an easy solution? Please contact me offline if you have any suggestions and other words of wisdom. Thanks in advance. Also, thanks for all the help you as a group have given me when I posted previous questions - sometimes I forget to say thank you.
}
} Paula.
}
} Paula Allan-Wojtas
} Research Scientist - Food Microstructure
} Agriculture and Agri-Food Canada
} Atlantic Food and Horticulture Research Centre
} Kentville, Nova Scotia Canada B4N 1J5
}
} Tel: (902) 679-5566
} FAX: (902) 679-2311
}
} email: allanwojtasp-at-em.agr.ca





From: Nestor J. Zaluzec :      zaluzec-at-sparc5.microscopy.com
Date: Tue, 20 Jun 2000 21:22:29 -0500
Subject: Administrivia: Microscopy Listserver Search Engine Now On-Line

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues....

A search engine for the Microscopy Listserver Archives is now on-line.

You can access it throught the Listserver Home Page at:

http://www.msa.microscopy.com/MicroscopyListserver

You can search the entire ~ 7 year history of the Listserver
if your heart desires...

Search options allow you to search the Subject, Author, and Message Text
seperately or together.

Please let me know of any problems, since I wrote the search engine I'll
have to fix it when it breaks.

It's not perfect, but hey it works....

Cheers...

Nestor
Your Friendly Neighborhood SysOp






From: Walck, Scott D. :      walck-at-ppg.com
Date: Tue, 20 Jun 2000 22:53:54 -0400
Subject: Speaking of User group meetings - How about Emispec?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Mansfield's reminded me that I would like to have some sort of users'
group meeting for Emispec users for mutual support and information exchange.
I have discussed this with the Emispec folks, but I don't think that they
have followed up on it. Could I seed some level of support for a
get-together at M&M MM so that I can send it to Emispec. Please respond to
me offline and I will send them the number and names of the people that
would desire something like that. I would like to see something in a
positive and instructive type of meeting.


-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)






From: Mark Blackford :      mgb-at-ansto.gov.au
Date: Wed, 21 Jun 2000 15:32:31 +1100
Subject: Rotory pump oil

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,

this is a message for EM people in Sydney or Woolongong, Australia.

The rotory pump on my JEOL 2020F TEM is short of oil and making a bit of
noise so I'm trying to find some quickly. Our JEOL service guys have
ordered some "MR100" but it may take a couple of weeks to arrive. Can any
one lend me a few hundred mls in the meantime?

Can anyone suggest a local supplier of this rotory pump oil?

Cheers,


Mark Blackford
TEM Group
Materials Division, ANSTO
PMB 1,
Menai, N.S.W.
Australia
2234

Phone 61 2 9717 3027
Fax 61 2 9543 7179

Disclaimer:
The views expressed in this E-mail message do not necessarily represent the
official views of ANSTO from which this message was conveyed.






From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Tue, 20 Jun 2000 17:49:44 +0100
Subject: Re: power problems with the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

My guess would be the current loading at start up. When you turn on
an SEM from cold, everything starts and draws current. In particular,
the rotary pump kicks in and draws a high current as it starts. You
will need to be able to set up the UPS so that it can handle this -
or set it on a short timer so that it automatically switches in, say,
10 mins after start up.

Regards,
--
Larry Stoter
JEOL (UK) Ltd
Silver Court, Watchmead, Welwyn Garden City, AL7 1LT, United Kingdom
tel: +44-(0)1707-377117, fax: +44-(0)1707-373254, e-mail: larrys-at-jeoleuro.com




From: Jo Verbeeck :      joverbee-at-ruca.ua.ac.be
Date: Wed, 21 Jun 2000 09:16:35 +0200 (METDST)
Subject: Re: power problems with the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

Probably you are having problems with the high switch on current the sem
draws. You could try switching on different parts of the SEM system not at
the same time (if possible). You could also consider to only power the
electronics of the SEM via the UPS, in that way you reduce the required
current drastically, while there is little use for a pump to sit on the
UPS power.
Alternatively you could look for a more powerfull UPS that can handle the
larger currents (for short moment is enough- see data sheets of different
UPS's).

Hope this helps...

Jo

*************************************************************
* Jo Verbeeck *
* University of Antwerp *
* Dept. EMAT (Electron Microscopy for Materials Research) *
* e-mail: joverbee-at-ruca.ua.ac.be *
* tel: +32(0)3 218 02 49 *
* fax: +32(0)3 218 02 57 *
*************************************************************

On Tue, 20 Jun 2000, Paula Allan-Wojtas wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi, all,
}
} I am posting this question to see if we can get some help for our building engineer to deal with problems we are having with our SEM (JEOL T330A).
}
} We suspect that we are having a problem with the building power supply to our SEM. Other folks in our building using other types of equipment have found it necessary to use power conditioners or UPS systems to run their equipment - these systems seem to do the trick for them.
}
} We are trying this approach with the power supply to our SEM, but find that there are some problems. When using the UPS system, we have to connect it up, put it on bypass, start up the SEM, then switch the UPS over after that - a real nuisance. If we don't do this, the SEM won't start.
}
} Does anyone else have this problem, and if so, has found an easy solution? Please contact me offline if you have any suggestions and other words of wisdom. Thanks in advance. Also, thanks for all the help you as a group have given me when I posted previous questions - sometimes I forget to say thank you.
}
} Paula.
}
} Paula Allan-Wojtas
} Research Scientist - Food Microstructure
} Agriculture and Agri-Food Canada
} Atlantic Food and Horticulture Research Centre
} Kentville, Nova Scotia Canada B4N 1J5
}
} Tel: (902) 679-5566
} FAX: (902) 679-2311
}
} email: allanwojtasp-at-em.agr.ca
}
}
}





From: Ron Doole :      ron.doole-at-materials.oxford.ac.uk
Date: Wed, 21 Jun 2000 08:36:37 +0100 (BST)
Subject: Re: EDS Window

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Terry,

If you are using water vapour in your environmental SEM you may
want to check the suitability of a Be window. I have a vague memory that
water will make holes in the thin (6-8um) Be, check it out with your
supplier.

I have used two polymer windows on an EDX detector in a TEM
gas reaction cell. The front window is on a replaceable mount in front of
the detector. This is easily exchangeable in case I cover it with reaction
products from the in-situ experiment; I don't want to see them in all
subsequent spectra. Much cheaper then a window replacement and can be
carried out by me on site. Contact me for further details if you want to.

Reducing the disturbance to the vacuum system gasses during
pumpdown and venting by limiting the speed may help prevent dust particles
damaging the window in the first place.

Regards,
Ron

On Tue, 20 Jun 2000 tellis2-at-hallmark.com-at-sparc5.microscopy.com wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} HI:
} For the second year in a row I am having to have our thin window (low
} element type-brand name with held) replaced.
} We have the EDS connected to a TOPCON SM-500 environmental SEM. I look at a
} lot industrial dusts for particle size and composition and the manuf. of
} the EDS system says they poke holes in the polymer window.
} Has anyone else have this problem?
} I have elected to have a thin beryllium installed this time since my
} boss is upset about spending $7000 every year plus the downtime and since
} low element detection is not critical .
} Thanks
} Terry Ellis
} Hallmark Cards Inc.
}
}
}

===========================================================================
Mr. Ron Doole e-mail ron.doole-at-materials.ox.ac.uk
Department of Materials, phone +44 (0) 1865 273701
University of Oxford, fax +44 (0) 1865 283333
Parks Road.
Oxford. OX1 3PH. UK.
============================================================================





From: Your Name :      KMG1TEX-at-bolton.ac.uk
Date: Wed, 21 Jun 2000 12:36:19 +0100
Subject: student requires information

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sir/Madam

I am currently studying for an MSc in Textile studies, at Bolton
Institute, England.

For part of my studies I have been analysing some polyester film that
has been treated with 10% sodium hydroxide under imposed load under
polarized light on an optical microscope. Vivid colours have been noted,
Reds, greens, and I am struggling to find information to outline what
these colours are actually indicating. I therefore write and ask for any
information you may deem relevant, I would be extremely grateful for.

I thank you for your time.

Yours Faithfully

Kelly Goodman

EMAIL: suite666-at-netscapeonline.co.uk





From: Leroux christine :      leroux-at-univ-tln.fr
Date: Wed, 21 Jun 2000 14:07:10 +0200
Subject: EDS analysis of Co in WC-Co

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
I am working on powders of mixed WC and Co and I have some problems with EDS.
EDS analysis on grains whose diffractions patterns are unambiguously
indexed in the Co structure give a majority of W, and EDS analysis on
grains whose diffractions patterns are unambiguously indexed in the WC
structure (without distortion and superstructures) always indicate the
presence of Co (with a majority of W). So I was wondering if there could be
any problem with EDS on Cobalt (because of magnetism????)

christine

**************************************
Christine leroux
Lab. M.M.I
Universite de Toulon-Var, Bat.R
B.P.132
F-83957 La Garde Cedex
tel: 00 33 (0) 4 94 14 25 07
fax: 00 33 (0) 4 94 14 21 68

**************************************





From: Robert.Carlton-at-aventis.com
Date: Wed, 21 Jun 2000 08:35:34 -0500
Subject: Re: EDS Window

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Terry,

I have an Electroscan 2010 and routinely have problems with the window. I
have ascribed it to particles flying around from the surface of the sample.
This generally occurs at startup when there may be some charging. We solved
the problem by retrofiting the detector so that it can be retracted from the
chamber when changing specimens. Its a pain in the butt, but less so than
having to replace windows.

Thanks
Robert Carlton
Aventis Pharmaceuticals
robert.carlton-at-aventis.com






From: Tobias Baskin :      BaskinT-at-missouri.edu
Date: Wed, 21 Jun 2000 09:06:04 -0500
Subject: Re: student ... (LM:polarized light)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
Yes, those colors are beautiful, arn't they?
The colors indicate the magnitude of birefringent retardation
in your sample. In essence, you are seeing subtraction colors
resulting from the diminution of intensity at certain wavelengths.
Say your sample has a retardation of 550 nm. When linearly polarized
light of that wavelength passes through the sample, it will be
retarded by exactly one wavelength, which is the same as zero
retardation. Thus, it will not be affected by the sample and will be
blocked by the analyzer. But light of longer or shorter wavelengths
will be retarded by more or less than a wave and so will emerge as
elliptically polarized light and thus will have a component
transmitted through the analyzer.
So for any actual sample retardation, the color will result
from exactly how much of each wavelength gets through. Because our
eyes are very sensitive to color, this has been used for more than
100 years to measure/estimate retardation. There is a chart that
reproduces the colors as a function of retardation. I have no idea if
this chart has made it on line but I would be careful. The colors are
very tricky to print and great care was used to get them right. You
would do far better to look this one up in your library. I am sure
the Bolton Textile Institute will have excellent books on polarized
light microscopy, and while they might be dusty, this particular
corner of science has been well understood for years and years.
Hope this helps,
Tobias Baskin



}
}
} Dear Sir/Madam
}
} I am currently studying for an MSc in Textile studies, at Bolton
} Institute, England.
}
} For part of my studies I have been analysing some polyester film that
} has been treated with 10% sodium hydroxide under imposed load under
} polarized light on an optical microscope. Vivid colours have been noted,
} Reds, greens, and I am struggling to find information to outline what
} these colours are actually indicating. I therefore write and ask for any
} information you may deem relevant, I would be extremely grateful for.
}
} I thank you for your time.
}
} Yours Faithfully
}
} Kelly Goodman
}
} EMAIL: suite666-at-netscapeonline.co.uk

_ ____ ^ __ ____ Tobias I. Baskin
/ \ / / \ / \ \ University of Missouri
/ | / / \ \ \ Biological Sciences
/___/ /__ /___ \ \ \__ 109 Tucker Hall
/ / / \ \ \ Columbia, MO 65211-7400 USA
/ / / \ \ \ voice: 573-882-0173
/ /____ / \ \__/ \____ fax: 573-882-0123




From: Sue Danielson :      sdaniels-at-post.its.mcw.edu
Date: Wed, 21 Jun 2000 10:11:03 -0500
Subject: TEM-micrographs for teaching manual

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everyone,

A colleage of mine here at the Medical College of Wisconsin is working on a
new teaching manual for our first-year medical students' cells & tissues
course. We are currently putting together a portion of the manual which
shows students electron micrographs of cells and organelles. We are looking
for high quality TEM images (non-copyrighted) of structures such as the
following:

Cytoskeleton (actin microfilaments, microvilli, microtubules, centrioles)

Cell Membrane (particularly intercellular junctions such as z. occludens,
adherens, desmosomes)

Protein synthesis & vescicles ( ribosomes, RER, golgi, endosomes,
lysosomes, peroxisomes, & vescicles-coated, endocytotic, secretory)

Nucleus (nuclear envelope, nuclear pores, chromatin, nucleolus)

MItochondria

We would be grateful to anyone who feels they have something useful to
contribute. The plan is to scan appropriate images for the lab manual and
return them as soon as possible to the owner. There will be a list at the
end of the manual acknowledging all contributors.

Thank you everyone,

Susan K. Danielson, MS
Neuromuscular Lab Coordinator
Dept. Neurology, Medical College of Wisconsin
ph: 414.259.3836
email: sdaniels-at-mcw.edu










From: Stephen Page :      stevep-at-micrometsys.com
Date: Wed, 21 Jun 2000 13:17:43 -0500
Subject: Job Posting-Technical Representative

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Here we grow again! We need more help.

We are looking for qualified professionals to provide applications-based
sales and technical support for our complete line of optical microscopes,
sample preparation equipment, consumables, and digital imaging systems in
the Dallas/Ft. Worth and Austin, TX areas. Must have at least 2 years
experience in failure analysis or materials analysis using microscopes and
sample preparation equipment. Prior sales experience is not necessary.
Understanding of dimensional measurement and image analysis systems a plus.

Please respond by e-mail to mms-at-micrometsys.com





From: mms :      mms-at-micrometsys.com
Date: Wed, 21 Jun 2000 13:21:38 -0500
Subject: Job Posting-Technical Representative

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Here we grow again! We need more help.

We are looking for qualified professionals to provide applications-based
sales and technical support for our complete line of optical microscopes,
sample preparation equipment, consumables, and digital imaging systems in
the Dallas/Ft. Worth and Austin, TX areas. Must have at least 2 years
experience in failure analysis or materials analysis using microscopes and
sample preparation equipment. Prior sales experience is not necessary.
Understanding of dimensional measurement and image analysis systems a plus.

Please respond by e-mail to mms-at-micrometsys.com





From: Gib Ahlstrand :      giba-at-puccini.cdl.umn.edu
Date: Wed, 21 Jun 2000 13:42:12 -0500 (CDT)
Subject: Re: acetone + LR Gold

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Responding to the message of
{3.0.5.32.20000620115226.008c7a80-at-mailserver.aecom.yu.edu}
from "David H. Hall" {hall-at-aecom.yu.edu} :

David,

I've done it, but the thing to watch out for is that if the temperature of your
acetone/LR Gold mixtures - say 1:1 acetone:LR Gold - drops below about -35 C,
some components of the LR Gold will start to freeze out, solutions gets cloudy.
For methanol sub-solution's, LR Gold starts to freeze out at even higher temps,
about -27 C.

Just experiment with your sub mixtures at various temperatures to see if you get
any freeze-out like this happening, before you do an actual sub run. If so just
make sure you stay warmer than freezing points.

I'm curious why this happens, anyone else got any info on this?

Gib Ahlstrand



} Does anyone have experience taking specimens from acetone into LR Gold
} resin? Some of the brochures on LR White seem to recommend against it, but
} we are hoping it may be OK for LR Gold. We are doing freeze substitution
} through acetone for EM-immunocytochemistry.
}
} Thanks in advance for your help.
} David H. Hall
} Center for C. elegans Anatomy
} Department of Neuroscience
} 1410 Pelham Parkway
} Albert Einstein College of Medicine
} Bronx, NY 10461
}
} phone (718) 430-2195 FAX (718) 430-8821
} hall-at-aecom.yu.edu
} website: www.aecom.yu.edu/wormem


Gib Ahlstrand
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN. USA. 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.cdl.umn.edu
http://biosci.umn.edu/MIC/consortium.html





From: Bob Price :      price-at-dcsmserver.med.sc.edu
Date: Wed, 21 Jun 2000 15:38:04 EDT
Subject: suggestions for symposia for M&M 2001 meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Even though we have not yet had our Microscopy and Microanalysis 2000
meeting in Philadelphia the Program Committee for the M&M 2001
meeting is preparing the symposia topics for the meeting in Long
Beach. If anyone has suggestions for topics that they would like to
see included in the Long Beach meeting please contact one of the
Program Committee Officers listed below. Please note that we may not
be able to respond to all suggestions for this year, but voicing your
opinion now may get a topic of interest on the program list for
upcoming years.

Thanks.

Bob Price

Program Officers:

Bob Price, M&M 2001 Program Chair
Price-at-med.sc.edu

Inga Holl Musselman, M&M 2001 MAS Program Co-chair
imusselm-at-utdallas.edu

Edgar Voelkl, M&M 2001 Program Vice-chair/M&M 2002 Program Chair
vog-at-ornl.gov

Robert L. Price
Director, Instrumentation
Resource Facility
USC School of Medicine
Garner's Ferry Road
Columbia, SC 29208
Phone: 803-733-3393
Fax:803-733-1533




From: jjerome-at-wfubmc.edu (Jay Jerome)
Date: Wed, 21 Jun 2000 17:07:19 -0400
Subject: Osmium disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I couldn't find this in the archives but I cannot imagine someone has
not asked this already. If so, please direct me to the approximate
month, year. If not, here goes:

We are decomissioning a laboratory refrigerator. Over the 30+ years of
use it has accumulated osmium black on the walls. It has been determined
by our EH&S people (after consultation with a hazardous waste shipping
compnay) that disposal requires a sealed, separate container for the
refrigerator, several licenses, and a hazardous waste truck. The
estimate is not in yet, but everyone is talking thousands of dollars. If
that is the only alternative for safe disposal, so be it. However, I
cannot imagine that there is not a safe alternative to decontaminate
before shipping. After all we clean up spills with corn oil and reduce
osmium tetroxide with 5% sodium bisulfite solution or sodium
metabisulfite. Any definitive advice or references will be
appreciated.

Thanks-

--
Jay
----------------------------------------------
- AKA: W. Gray Jerome, Ph.D. -
- Department of Pathology -
- Wake Forest University School of Medicine -
- Winston-Salem, NC 27157-1092 -
- Ph: 336-716-4972, 336-716-2675 -
- Fax: 336-716-6174 -
- E-mail: jjerome-at-wfubmc.edu -
----------------------------------------------






From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Wed, 21 Jun 2000 19:56:33 -0500
Subject: Refrigerator osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

W. Gray Jerome wrote:
===========================================================
I couldn't find this in the archives but I cannot imagine someone has not
asked this already. If so, please direct me to the approximate month, year.
If not, here goes:

We are decomissioning a laboratory refrigerator. Over the 30+ years of use
it has accumulated osmium black on the walls. It has been determined by our
EH&S people (after consultation with a hazardous waste shipping compnay)
that disposal requires a sealed, separate container for the refrigerator,
several licenses, and a hazardous waste truck. The estimate is not in yet,
but everyone is talking thousands of dollars. If that is the only
alternative for safe disposal, so be it. However, I cannot imagine that
there is not a safe alternative to decontaminate before shipping. After all
we clean up spills with corn oil and reduce osmium tetroxide with 5% sodium
bisulfite solution or sodium metabisulfite. Any definitive advice or
references will be appreciated.
=============================================================
I will risk showing what maybe I don't know, but this is really an important
question. Laboratories should not be wasting thousands of dollars
needlessly, when other alternatives are available.

I thought that Os (IV) oxide, that is, the tetroxide form of osmium, was
clear, after all, crystals in ampoules are fairly clear or translucent, and
4% aqueous osmium tetroxide is water clear. And I thought that in the
reduced form, that is, the dioxide or Os (II) form, the color was black.
Putting it another way, if it is in the reduced form (e.g. black), it is not
in its "hazardous" form, and indeed might even be fairly innocuous. A
**quick** survey of shipping regulations does not even indicate that osmium
(II) oxide is a controlled material from the standpoint of shipping. I want
to emphasize the word "quick", but I could not find a UN number asigned for
it.

If that is the case, and if I am right, why could not the black deposit be
collected, perhaps by way of some rubbing and scrubbing, and disposed of
(or better yet, recycled) and the refrigerator disposed of as any ordinary
refrigerator might be dealt with?

Remember, I am not saying that this could be done, legally or otherwise, I
am just asking what it is that I might be missing here.

Disclaimer: SPI Supplies is a major supplier of osmium tetroxide to EM
laboratories worldwide.

Chuck
============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
============================================






From: Paul-Laurent Capron Jr. :      Paul-Laurent.capron.jr-at-laborimpex.be
Date: Wed, 21 Jun 2000 19:07:12 -0500
Subject: chemical reference standards for the analysis of a Ni-base alloy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



CAN SOMEBODY HELP ME?:

-We need chemical reference standards for the analysis of a Ni-base alloy.

-We intend to use the energy-dispersive X-ray analysis, coupled to a
Scanning
Electron Microscope(SEM).

-The composition of the alloy is going to vary between 18 to 28 wt%
Aluminium ;
Al is the most important element to be tested, the remainder of the alloy
being
mostly Ni of course.

-Secondary main elements are : Cr from 4 to 10 wt% ; Co from 10 to 15
wt% ; Elements in low concentration (below 3%) are Ti and Mo.

-Accuracy : 0.5 % in Al is required.


Questions are :

-How many different standards would we need ?

-What should be the chemical variations from standard to standard to reach
the
required accuracy ?

-What would be the price for such standards ? Time for delivery ? Where to
order this ?

-Can you send us information (reprints from technical journals, articles) on
the
topic of chemical analysis by means of energy-dispersive (or
wavelength-dispersive) X-ray systems ?

If you need any further information, please contact me.

In anticipation, many thanks for a prompt answer.

Yours faithfully,

PAUL-LAURENT CAPRON JR.
S.A. LABORIMPEX N.V.
RUE DES ALLIES 78-80
BONDGENOTENSTRAAT 78-80
1190 BRUSSEL/BRUXELLES
BELGIUM
TEL.: 00 32 2 345.99.94
FAX.: 00 32 2 347.39.63






From: Barry Searle :      B.Searle-at-unsw.edu.au
Date: Wed, 21 Jun 2000 19:23:37 -0500
Subject: JEOL Coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fellow EM Users, Just a short query ----- would anyone installed a
modern day alternative to the 1920's !!! style glass ionization vacuum
gauge ("electron current") + electronics box for the JEOL 4B coater?
Thankyou for your time, Barry EM UNIT UNSW






From: jim :      jim-at-proscitech.com.au
Date: Thu, 22 Jun 2000 11:49:58 +1000
Subject: RE: Osmium disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Very funny, in every meaning of that word. Of course, the waste company would
make a meal of it, why ask?

Its not Os04 that is coating the inside of the fridge, but a tiny amount of
osmium metal. The metal is not particularly toxic. Os and Ir are the heaviest
metals, with a specific gravity over 22, but nonetheless, its hard to see that
you would have 2g of Os in any refrigerator coating. Finely dispersed Os
produces tiny amounts of Os04 - if that is a problem, than that would have been
so in the confined lab space. If that fridge was dumped conventionally in a
land-fill, the tiny amount of OsO4 would be reduced before ever reaching the
surface. If it was smelted for scrap, arguably it would, however, immeasurably
improve the quality of the scrap.
There is an argument for smearing a bit of vegetable oil over the inside of the
fridge prior to disposal.
Have those safety people given their fully developed reasons? I love to know
them.

The bigger picture is interesting too. Safety people are not selected for, nor
seem to get taught much logic or a sense of proportion. So we seem to get more
and more hassles about things that hardly matter at all, but there is no money
and often no objection to the really large environmental issues. Its been said
that the old Roman civilasation's collapse was partially due to the difficult
Roman numerals, which made the efficient administration of the empire
impossible. Our "Western Society" has seen a rapid increase of useless, if not
counterproductive rules and regulations, I fear if that trend continues that
these "well-meaning" rules will undermine our society.
Cheers
Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Thursday, June 22, 2000 7:07 AM, Jay Jerome [SMTP:jjerome-at-wfubmc.edu] wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I couldn't find this in the archives but I cannot imagine someone has
} not asked this already. If so, please direct me to the approximate
} month, year. If not, here goes:
}
} We are decomissioning a laboratory refrigerator. Over the 30+ years of
} use it has accumulated osmium black on the walls. It has been determined
} by our EH&S people (after consultation with a hazardous waste shipping
} compnay) that disposal requires a sealed, separate container for the
} refrigerator, several licenses, and a hazardous waste truck. The
} estimate is not in yet, but everyone is talking thousands of dollars. If
} that is the only alternative for safe disposal, so be it. However, I
} cannot imagine that there is not a safe alternative to decontaminate
} before shipping. After all we clean up spills with corn oil and reduce
} osmium tetroxide with 5% sodium bisulfite solution or sodium
} metabisulfite. Any definitive advice or references will be
} appreciated.
}
} Thanks-
}
} --
} Jay
} ----------------------------------------------
} - AKA: W. Gray Jerome, Ph.D. -
} - Department of Pathology -
} - Wake Forest University School of Medicine -
} - Winston-Salem, NC 27157-1092 -
} - Ph: 336-716-4972, 336-716-2675 -
} - Fax: 336-716-6174 -
} - E-mail: jjerome-at-wfubmc.edu -
} ----------------------------------------------
}
}





From: jim :      jim-at-proscitech.com.au
Date: Thu, 22 Jun 2000 12:01:59 +1000
Subject: RE: Rotory pump oil

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rotary pump oil is a refined mineral oil, which needs to have low vapour
pressure, low viscosity and good lubricating properties. such oils are made by
"oil companies". In my lab days I never had the luxury to purchase oil from a
microscope manufacturer, because their on-cost for such items are horrific and
real and my budgets were small.

Disclaimer: ProSciTech sell Rotary Pump Oil. Don't even ask to ship it to USA;
the material is cheap, but expensive to ship. Our price includes shipping
within Australia only.
Cheers
Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Wednesday, June 21, 2000 2:33 PM, Mark Blackford [SMTP:mgb-at-ansto.gov.au]
wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear all,
}
} this is a message for EM people in Sydney or Woolongong, Australia.
}
} The rotory pump on my JEOL 2020F TEM is short of oil and making a bit of
} noise so I'm trying to find some quickly. Our JEOL service guys have
} ordered some "MR100" but it may take a couple of weeks to arrive. Can any
} one lend me a few hundred mls in the meantime?
}
} Can anyone suggest a local supplier of this rotory pump oil?
}
} Cheers,
}
}
} Mark Blackford
} TEM Group
} Materials Division, ANSTO
} PMB 1,
} Menai, N.S.W.
} Australia
} 2234
}
} Phone 61 2 9717 3027
} Fax 61 2 9543 7179
}
} Disclaimer:
} The views expressed in this E-mail message do not necessarily represent the
} official views of ANSTO from which this message was conveyed.
}
}





From: jim :      jim-at-proscitech.com.au
Date: Thu, 22 Jun 2000 15:20:45 +1000
Subject: RE: JEOL Coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I remember using the old Geissler tube on an evaporator 32 years ago and very
likely they existed long before then. They were actually a bit of an ongoing
maintenance expense, but I loved the pretty colours. Cannot see any reasons why
other type gauges could not be used, except the manufacturer would find the
Geissler tube a cheaper alternative.

Disclaimer: ProSciTech supplies vacuum gauges, circa year 2000

Cheers
Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Thursday, June 22, 2000 10:24 AM, Barry Searle [SMTP:B.Searle-at-unsw.edu.au]
wrote:
}
} Fellow EM Users, Just a short query ----- would anyone installed a
} modern day alternative to the 1920's !!! style glass ionization vacuum
} gauge ("electron current") + electronics box for the JEOL 4B coater?
} Thankyou for your time, Barry EM UNIT UNSW
}




From: Robert H. Olley :      r.h.olley-at-reading.ac.uk
Date: Thu, 22 Jun 2000 09:20:35 +0100 (BST)
Subject: Re: student requires information

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



to: Kelly Goodman,

Greetings!

Regarding the colours in the PET film, the optical part is as Tobias
Baskin says. In regard to the polymer science side of it, sretched films
or fibres show optical anisotropy, because the polarizability along the
polymer chain is different from that transverse to it. The more highly
stretched the fibre, the more the chains will be oriented, and this will
increase the optical anisotropy of the fibre. Multiply this by the
thickness of the fibre and that will give you the birefringence. Most
likely you will be seeing coloured fringes parallel to the fibre
direction, which indicate the different viewing thickness through the
different parts of the fibre, just like the colours in an oil slick under
a car indicate different thicknesses of an oil film.

+------------------------------------------------------------------------+
| Robert H.Olley Phone: |
| J.J.Thomson Physical Laboratory {direct line +44 (0) 118 9318572 |
| University of Reading {University internal extension 7867 |
| Whiteknights Fax +44 (0) 118 9750203 |
| Reading RG6 6AF Email: R.H.Olley-at-reading.ac.uk |
| England URL: http://www.reading.ac.uk/~spsolley |
+------------------------------------------------------------------------+










From: Koh YinHsian-CYK006 :      Yinhsian.Koh-at-Motorola.com
Date: Thu, 22 Jun 2000 18:37:53 +0800
Subject: Interface Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I am not familiar with microscopy. The most I know is how SEM/EDX can
examine a cross sectional sample.

Currently, I am looking for some methodology to analyse interface structure,
elementry and chemical compound especially between the solder and metal,
i.e. the IMC layer. The solder height will be ranging from 2-4 mil (SMT
process). Wondering anybody can advice the best method to analyse the IMC
layer. EDX is capable of detecting the element contain insde the IMC but I
am seeing more and more alloy up to ternary or quartenary alloy at the
layer. Problem is I do not have any idea about where and how it come from.

Appreciate your help.

Thank you.

Regards,
YH Koh




From: Kristian Ukkonen :      kukkonen-at-cc.hut.fi
Date: Thu, 22 Jun 2000 13:54:52 +0300 (EET DST)
Subject: JVG-N2 vacuum gauge and forgel tubes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello,

I'm looking for a manual for the JVG-N2 ionization
vacuum gauge. I've already tried JEOL and they could
find JVG-N1 manual (thanks!), but not -N2, and -N1 and
-N2 are different enough (different connectors, apparently
EB grid heating in -N2 etc.) that the -N1 manual is not
enough to get the gauge operational.

So, anyone with manual for JVG-N2 ??

btw: The JVG-N1 manual mentions "Forgel tube" which looks
in drawing just like normal hot-filament ionization
gauge tube.. Is "Forgel tube" just an old term for one?

Thanks,

Kristian Ukkonen.
kukkonen-at-cc.hut.fi





From: White, Woody N :      Woody.N.White-at-mcdermott.com
Date: Thu, 22 Jun 2000 08:50:00 -0500
Subject: Re:Interface Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




For some idea of what can be done, check out
the soldering iron tip images and x-ray
data at my web site.

Keep in mind that I did not not take time
to achieve the best possible polish and
most of the work was at rather low magnification.

Site: http://woody.white.home.att.net

Woody White

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hello,

I am not familiar with microscopy. The most I know is how SEM/EDX can
examine a cross sectional sample.

Currently, I am looking for some methodology to analyse interface structure,
elementry and chemical compound especially between the solder and metal,
i.e. the IMC layer. The solder height will be ranging from 2-4 mil (SMT
process). Wondering anybody can advice the best method to analyse the IMC
layer. EDX is capable of detecting the element contain insde the IMC but I
am seeing more and more alloy up to ternary or quartenary alloy at the
layer. Problem is I do not have any idea about where and how it come from.

Appreciate your help.

Thank you.

Regards,
YH Koh




From: Donald Lovett :      lovett-at-tcnj.edu
Date: Thu, 22 Jun 2000 09:06:47 -0400 (EDT)
Subject: Re: Refrigerator osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Wed, 21 Jun 2000, Garber, Charles A. wrote:

} question. Laboratories should not be wasting thousands of dollars
} needlessly, when other alternatives are available.
}
} I thought that Os (IV) oxide, that is, the tetroxide form of osmium, was
} clear, after all, crystals in ampoules are fairly clear or translucent, and
} 4% aqueous osmium tetroxide is water clear. And I thought that in the
} reduced form, that is, the dioxide or Os (II) form, the color was black.
} Putting it another way, if it is in the reduced form (e.g. black), it is not
} in its "hazardous" form, and indeed might even be fairly innocuous. A
} **quick** survey of shipping regulations does not even indicate that osmium
} (II) oxide is a controlled material from the standpoint of shipping. I want
} to emphasize the word "quick", but I could not find a UN number asigned for
} it.
}
} If that is the case, and if I am right, why could not the black deposit be
} collected, perhaps by way of some rubbing and scrubbing, and disposed of
} (or better yet, recycled) and the refrigerator disposed of as any ordinary
} refrigerator might be dealt with?
}

The problem is that any form of osmium is a heavy metal, and therefore
should be treated as one would cadmium or arsenic. It is not innocuous.

Don
}

______________________________________________________________________
Donald L. Lovett e-mail: lovett-at-tcnj.edu
Assoc. Professor, Dept. of Biology voice: (609) 771-2876
P.O. Box 7718 fax: (609) 637-5118
The College of New Jersey
Ewing, NJ 08628-0718







From: yliu-at-unlserve.unl.edu (Yi Liu)
Date: Thu, 22 Jun 2000 08:38:21 -0500
Subject: Need info on magnetic structure characterization

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Colleagues:

We have made magnetic nanodots and metastable magnetic phases. The
physical structures are well revealed by TEM. However, we would like to
relate the magnetic properties with magnetic structure and physical
structure. Are there other methods such as neutron diffraction, X-ray or
spin polarized electrons that can characterize the magnetic
structure/ordering. Please contact me if you could offer such
info/collaboration. Thanks a lot.

Yi

*******************************************************************
Yi Liu
Department of Mechanical Eng. and CMRA
104 N Walter Scott Engineering Center
University of Nebraska-Lincoln
Lincoln, NE 68588-0656
Tel. (402) 472-7759 (Office)
Tel. (402) 472-8762 (EM lab)
Fax (402) 472-1465
Email: yliu-at-unlserve.unl.edu
*******************************************************************






From: Anthony Garratt-Reed :      tonygr-at-mit.edu
Date: Thu, 22 Jun 2000 10:00:13 -0400
Subject: Re: Interface Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


YinHsian -

This is a very complex problem. You don't mention the thickness of the
intermetallic layers. EDX analysis has relatively poor spatial resolution
in the SEM (of the order of microns) so you can clearly see features that
you cannot analyze well. People often use backscatter electron imaging
(which can be calibrated) in cases such as yours. You should read a good
book on SEM - the subject is too complex to cover in an e-mail like this.

If you can make thin foils of the sample, AEM (analytical electron
microscopy) analysis would overcome most of the problems. A LaB6
instrument will give you better than 10 nm analytical resolution, while a
field-emitter will approach 1 nm resolution. The problem, though, will be
specimen preparation. I would be tempted to suggest using a microtome to
cut thin slices, but I don't know the form of your material.

Good luck!

Tony Garratt-Reed


At 06:37 PM 06/22/2000 +0800, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6479
**






From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Thu, 22 Jun 2000 09:02:05 -0700
Subject: Re: chemical reference standards for the analysis of a Ni-base

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Paul-Laurent,
The usual standards that are used for alloy analysis in the SEM+EDX are pure
element standards. You can prepare pure (three or four nines, e.g. 99.9 or
99.99%) metal standards yourself, or any of the EM catalog houses will list
standards consisting of pure metals or minerals covering the range of most
common elements. They are fairly expensive (~$2000US) and I am not sure of
delivery time. Tousimis is one standard supplier that I know of. The same
standards are used for EDX or WDX analysis. You collect the spectra from
these standards under carefully reproducible conditions in your SEM and
follow the instructions in your EDX software to do fully standard analysis
with matrix corrections. You can store the standard spectra in the EDX
computer, so you only need to run them once for each set of conditions.
It is quite difficult to get an accurate and precise analysis of Al in Ni
alloys, because there is a large correction factor for absorption of the Al
x-rays by the Ni matrix, and because the best SEM conditions for Ni
analysis, 20 kV, is a large overvoltage for Al. It is very helpful if you
can make or obtain a well-characterized, homogeneous alloy in the range of
the alloy you will be testing, so you can check the results of your EDX
analysis. The last time I did a careful phase analysis of an alloy
consisting of 5% Al in Zn, I eventually went to analysing the Al at 10 kV
and doing the Zn by difference. This was the only way to get the Al results
with the accuracy required and agreeing with the phase diagram. With the
other elements present in your system, you cannot use the
element-by-difference method, but you could check to see if you get the same
Al results at 10 kV, doing the Ni by difference (forgetting the other
elements), as you get in your full, standard analysis at 20 kV.
Most of the EDX supplier companies have excellent instruction books that
cover the topics of chemical analysis by EDX and the exact conditions
required for accurate analysis. Some offer courses in EDX analysis and they
are a very good way to get familiar with the strengths and weaknesses of the
system.
Please contact me if you need any more information.
At 07:07 PM 6/21/00 -0500, you wrote:

} CAN SOMEBODY HELP ME?:
}
} -We need chemical reference standards for the analysis of a Ni-base alloy.
}
} -We intend to use the energy-dispersive X-ray analysis, coupled to a
} Scanning
} Electron Microscope(SEM).
}
} -The composition of the alloy is going to vary between 18 to 28 wt%
} Aluminium ;
} Al is the most important element to be tested, the remainder of the alloy
} being
} mostly Ni of course.
}
} -Secondary main elements are : Cr from 4 to 10 wt% ; Co from 10 to 15
} wt% ; Elements in low concentration (below 3%) are Ti and Mo.
}
} -Accuracy : 0.5 % in Al is required.
}
}
} Questions are :
}
} -How many different standards would we need ?
}
} -What should be the chemical variations from standard to standard to reach
} the
} required accuracy ?
}
} -What would be the price for such standards ? Time for delivery ? Where to
} order this ?
}
} -Can you send us information (reprints from technical journals, articles) on
} the
} topic of chemical analysis by means of energy-dispersive (or
} wavelength-dispersive) X-ray systems ?
}
} If you need any further information, please contact me.
}
} In anticipation, many thanks for a prompt answer.
}
} Yours faithfully,
}
} PAUL-LAURENT CAPRON JR.

Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca





From: Tom Januszewski :      tom.januszewski-at-email.swmed.edu
Date: Thu, 22 Jun 2000 12:57:41 -0500
Subject: Critical Point Dryers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Listservers,
We are in the process of replacing our old
critical point dryer and have narrowed our
selection to the following choices: Tousimis
Samdri 795 semi automatic and the EMS 850
critical point dryers.
If anyone out there is familiar with either
of these instruments and would care to
comment on their advantages and
disadvantages, I would appreciate hearing
from you. Offline replies are welcome. I
will report back with summary of results.

Tom Januszewski
Senior Electron Microscopist
Molecular and Cellular Imaging Facility
UT Southwestern Medical Center at Dallas
5323 Harry Hines Blvd.
Dallas, TX 75390-9039
Phone: 214-648-7291
Fax: 214-648-6408
Email: tom.januszewski-at-email.swmed.edu





From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Thu, 22 Jun 2000 13:37:42 -0700
Subject: Re: JVG-N2 vacuum gauge and forgel tubes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Kristian,
I also have an old JEOL evaporator (JEE-4B) and, yes, the Forgel tube was
just a hot-wire ionization gauge. I got tired of buying new ones, so I
successfullly replaced that with a cold cathode gauge attached to a tube of the
right diameter.
At 01:54 PM 6/22/00 +0300, you wrote:
} Hello,
}
} I'm looking for a manual for the JVG-N2 ionization
} vacuum gauge. I've already tried JEOL and they could
} find JVG-N1 manual (thanks!), but not -N2, and -N1 and
} -N2 are different enough (different connectors, apparently
} EB grid heating in -N2 etc.) that the -N1 manual is not
} enough to get the gauge operational.
}
} So, anyone with manual for JVG-N2 ??
}
} btw: The JVG-N1 manual mentions "Forgel tube" which looks
} in drawing just like normal hot-filament ionization
} gauge tube.. Is "Forgel tube" just an old term for one?
}
} Thanks,
}
} Kristian Ukkonen.
Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca





From: Chris Jeffree :      cjeffree-at-srv0.bio.ed.ac.uk
Date: Thu, 22 Jun 2000 21:57:57 +0100
Subject: Re: Refrigerator osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This doesn't seem to cover it. Gold and platinum are heavy metals.
So are we to treat them as we treat osmium tetroxide?
Where does light end and heavy begin anyway? There is a real
need here for a proper definition of the risk.

}
} The problem is that any form of osmium is a heavy metal, and therefore
} should be treated as one would cadmium or arsenic. It is not innocuous.
}
} Don
} }
}
} ______________________________________________________________________
} Donald L. Lovett e-mail: lovett-at-tcnj.edu
} Assoc. Professor, Dept. of Biology voice: (609) 771-2876
} P.O. Box 7718 fax: (609) 637-5118
} The College of New Jersey
} Ewing, NJ 08628-0718
}
}
}
}


~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Dr Chris Jeffree
University of Edinburgh
Biological Sciences EM Facility
Daniel Rutherford Building
King's Buildings EDINBURGH EH9 3JH
Tel: +44 (0) 131 650 5345
FAX: +44 (0) 131 650 6563

Inveresk Cottage, 26 Carberry Road,
Inveresk, Musselburgh, Midlothian EH21 8PR, UK
Tel. +44 (0) 131 665 6062 / Mobile 0410 585 401
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~




From: Corazon Bucana :      bucana-at-audumla.mdacc.tmc.edu
Date: Thu, 22 Jun 2000 16:18:36 -0500
Subject: EM autoradiography

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I received a request for EM autoradiography of C14 labeled polymers. The
idea is to trace the distribution of the polymer once it is internalized in
the cell. It has been a long while since I have done autoradiography and I
was wondering if there are new books and information that is available out
there. Is anyone still using this technique. What are the alternatives? I
am not too keen on handling radioactive materials again. I would appreciate
comments.

*******************************************************
Corazon D. Bucana, Ph.D.
Department of Cancer Biology
U.T. M.D. Anderson Cancer Center
1515 Holcombe Blvd. Box 173
Houston, Texas 77030
Phone: (713) 792-8106
FAX: (713) 792-8747
Email:bucana-at-audumla.mdacc.tmc.edu
FAX: (713) 792-8747





From: Ken Long :      long-at-ROBLES.CALLUTHERAN.EDU
Date: Thu, 22 Jun 2000 15:05:00 -0700
Subject: undergrad SEM use

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings:
I am a new subscriber to this list and am looking for input/advice
regarding the use of SEM's in undergraduate research. I teach at a liberal
arts university with only undergraduate programs in the natural sciences.
A geology colleague and myself (I'm a biologist) are exploring possible
funding sources which would allow us to purchase an SEM that would be used
by biology and geology undergrads in their independent and senior thesis
projects. (We currently have a non-functioning ETEC that we would like to
replace -- we've been advised that further repair and head-banging would
not be cost effective). If anyone is currently engaged in SEM-based
research with undergrads, I would like to get your feedback on your
experiences (both the pluses and minuses). Although we don't have a TEM,
input from those using TEM with undergrads would also be helpful.

Specific Questions:

1) What types of projects are your students pursuing?

2) Are you conducting interdisciplinary projects? e.g., micropaleontology
or environmental applications.

3) Are there particular models of SEMs which would be more user friendly
since we would be training undergrads and we would also be the primary
trainers and technicians?

4) Have you developed an undergrad EM course?

Thanks,
Ken

-----------------------------------------------------------------------
Kenneth Long, PhD email: long-at-clunet.edu
Associate Professor, Biology Office: (805) 493-3346
California Lutheran University Fax: (805) 493-3392
60 West Olsen Rd. #3700 http://www.clunet.edu/Biology/Anatomy
Thousand Oaks, CA 91360






From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 22 Jun 2000 17:11:55 -0500
Subject: M&M-Facility Management

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Fellow Microscopists:
Stuart Mckernan,M&M'00 Program Chair, has arranged for an additional "EXPERTS" session for the topic of Facility Management. This special session will be held on Wednesday, August 16 at 9:00Am in Room 106. Advertising will be done by flyers, daily news bulletin, and word of mouth.

The room will be available all morning so that we can be somewhat flexible in timing. However, in order to try to make the session as constructive as possible, I would like to identify 2-3 main topics of interest to lab managers from both academic and industrial facilities. These topics will be covered first. Additional topics can then be discussed based on interest and time.

I would appreciate it if those who are interested in attending this session would provide the following:
a) 2-3 topics of primary interest to you ranked in order of that interest.
b) suggestions of individuals who might be asked to introduce a specific topic so as to give a framework for further discussion.

I am heading out for vacation on July 1 and will unsubscribe from the list a few days prior to then. Please send the above information directly to me rather than to the list so I am sure to get it. We will announce the topics to be discussed and approximate times they will be discussed so that you can choose when you want to attend the session.

If the session is well attended, then it may be added to the agendas of future meeting. Hope to hear from many of you regarding the topics.
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057





From: E. J. McKenzie :      elizm-at-pdx.edu
Date: Thu, 22 Jun 2000 15:50:11 -0700
Subject: Formaldehyde problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) id SAA15867
for dist-Microscopy; Thu, 22 Jun 2000 18:00:09 -0500 (CDT)
Received: from no_more_spam.com (sparc5 [206.69.208.10])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with SMTP id RAA15863
for "MicroscopyFilteredEmail-at-msa.microscopy.com"; Thu, 22 Jun 2000 17:59:39 -0500 (CDT)
Received: from pdx.edu (tuttle.oit.pdx.edu [131.252.120.29])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with ESMTP id RAA15856
for {microscopy-at-sparc5.microscopy.com} ; Thu, 22 Jun 2000 17:59:28 -0500 (CDT)
Received: from [131.252.192.235] (host-192-235.dhcp.pdx.edu [131.252.192.235])
by pdx.edu (8.9.3/8.9.3) with ESMTP id PAA26977
for {microscopy-at-sparc5.microscopy.com} ; Thu, 22 Jun 2000 15:53:53 -0700 (PDT)
Message-Id: {200006222253.PAA26977-at-pdx.edu}
X-Mailer: Microsoft Outlook Express Macintosh Edition - 4.5 (0410)


Hi everyone,

I'm currently making 1.5 litre batches of 4% formaldehyde
using paraformaldehyde, sodium hydroxide and distilled water and not
having much success. The mixture immediately begins to polymerize. The
recipe I used is simply an amplified version of Karnovsky's recipe for 25ml
of 4% (which works perfectly). I am wondering whether my assumption of a
simple linear relationship between the proportions of NaOH and
paraformaldehyde is incorrect. Making stock solutions of higher
concentrations
does not work either.

Does anyone know a more successful method of preparing large
volumes of formaldehyde? We do not want to use phosphates or
preservatives.

regards
Elizabeth McKenzie
--------------------------------------------------
Geomicrobiology and Electron Microscopy Laboratory
Room S9 Cramer Hall
1721 SW Broadway
Portland State University
Portland
OR97201

ph:503 725 3362
fax:503 725 3025





From: Barry Searle :      B.Searle-at-unsw.edu.au
Date: Thu, 22 Jun 2000 21:10:45 -0500
Subject: JEE 4B/4C Coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fellow EM Users, Would anyone have a list of the O rings for use on
the JEOL JEE4B/4C vacuum evaporator. The photocopy of pages from the
manual that I have does not list any o-rings. Thankyou Barry EM UNIT
UNSW






From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Thu, 22 Jun 2000 21:07:37 -0700
Subject: Fwd: Polaron E5200 control board

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


One response. No follow through. Board is going
into the circular file.

gg



} Date: Thu, 15 Jun 2000 07:55:38 -0700
} To: MSA listserver
} From: "Dr. Gary Gaugler" {gary-at-gaugler.com}
} Subject: Polaron E5200 control board
}
} I have a control PC board for the E5200 sputter coater.
} This is the model with an Intel single chip MPU on one
} end and a 4-conductor socket on the other end. The
} board uses a VME connector for main interface.
}
} Coater is trashed. Board is OK. If anybody can use
} the board, first request gets it.
}
} gary g.





From: jim :      jim-at-proscitech.com.au
Date: Fri, 23 Jun 2000 14:48:14 +1000
Subject: RE: Refrigerator osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I had a look at the excellent site
http://www.webelements.com/index.html

clicking on osmium and then asking for "biology" this info is given:
"Osmium metal does not normally cause problems as it is relatively unreactive
but all osmium compounds should be regarded as highly toxic. The metal dust is
an irritant and presents a fire and explosion hazard. Osmium oxide, OsO4, is
highly toxic, and boils at 130?C (760 mm). Concentrations in air as low as 10-7
g m-3 can cause lung congestion, skin damage, and severe eye damage. The oxide,
in particular, should only ever be handled by a properly qualified chemist."
We all know about the tetroxide, but the point I made in my previous email was
that the fridge contamination is Osmium metal and this has little reactivity.
Osmium when finely dispersed would produce a tiny amount of tetroxide. Possibly
not enough to worry while in the lab, certainly not if its in landfill or
smelted.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Friday, June 23, 2000 6:58 AM, Chris Jeffree
[SMTP:cjeffree-at-srv0.bio.ed.ac.uk] wrote:}
}
} This doesn't seem to cover it. Gold and platinum are heavy metals.
} So are we to treat them as we treat osmium tetroxide?
} Where does light end and heavy begin anyway? There is a real
} need here for a proper definition of the risk.
}
} }
} } The problem is that any form of osmium is a heavy metal, and therefore
} } should be treated as one would cadmium or arsenic. It is not innocuous.
} }
} } Don
} } }
} }
} } ______________________________________________________________________
} } Donald L. Lovett e-mail: lovett-at-tcnj.edu
} } Assoc. Professor, Dept. of Biology voice: (609) 771-2876
} } P.O. Box 7718 fax: (609) 637-5118
} } The College of New Jersey
} } Ewing, NJ 08628-0718
} }
} }
} }
} }
}
}
} ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
} Dr Chris Jeffree
} University of Edinburgh
} Biological Sciences EM Facility
} Daniel Rutherford Building
} King's Buildings EDINBURGH EH9 3JH
} Tel: +44 (0) 131 650 5345
} FAX: +44 (0) 131 650 6563
}
} Inveresk Cottage, 26 Carberry Road,
} Inveresk, Musselburgh, Midlothian EH21 8PR, UK
} Tel. +44 (0) 131 665 6062 / Mobile 0410 585 401
} ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: Oldrich Benada :      benada-at-biomed.cas.cz
Date: Fri, 23 Jun 2000 11:11:15 +0200
Subject: A hole in the EDX spectrum II

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear colleagues,
Many thanks for all responses to my question considering the hole in
the EDX spectrum.
We had a service engineer in our lab and he tried to solve the
problem but he was not successful.
I have prepared WEB page with the images of spectra, that were
recorded from following samples:
Cu_Al sample (usually used for calibration)
Ti grid
AL grid
K standard

The url of that page is following:
http://www.biomed.cas.cz/~benada/EDX_page/index.htm

Please, if anyone would be so kind and could comment the spectra, we
will be very happy.
Best regards from Prague
Oldrich Benada


+-----------------------------------+
Oldrich Benada
Acad. Sci. CR
Institute of Microbiology
Laboratory of electron microscopy
Videnska 1083
CZ - 142 20 Prague 4 - Krc
Czech Republic
+------------------------------------+
Phone: +420-2-4752399
Fax: +420-2-4715743
WEB:http://www.biomed.cas.cz/mbu/lem113/lem.htm




From: Bart De Pauw :      Bart.DePauw-at-rug.ac.be
Date: Fri, 23 Jun 2000 11:42:12 +0200
Subject: Re: Formaldehyde problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There's nothing mysterious about this. It should be possible to
make any volume of formaldehyde solution over a wide range of
concentrations. We routinely make 10% solutions without
problems in volumes much greater than 25ml. There is no reason
to suppose that the volume is relevant, and the concentration is not
critical.
The temperature needs to be } 60oC and the solution needs to be
slightly alkaline. Otherwise it is extremely straightforward.
Could you tell us your procedure, step by step? Then maybe it will
be possible to diagnose the problem.
Chris


Date sent: Thu, 22 Jun 2000 15:50:11 -0700


Commerially available formaldehyde contains formic acid and methanol and is
therefore unsuitable for EM. Instead, one usually prepares a solution of
formaldehyde from paraformaldehyde. An aqueous solution (40%, w/v) can be made
by warming for 1 h at 65ˇV (constant stirring). Milkyness can be overcome by
adding a few drops of a 40% (w/v) NaOH solution. This solution is stable for
several weeks at 4ˇc (in the dark).
Hopes this helps.

De Pauw Bart
Ghent University
Faculty of Veterinary Medicine
Department Morphology


"E. J. McKenzie" schreef:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hi everyone,
}
} I'm currently making 1.5 litre batches of 4% formaldehyde
} using paraformaldehyde, sodium hydroxide and distilled water and not
} having much success. The mixture immediately begins to polymerize. The
} recipe I used is simply an amplified version of Karnovsky's recipe for 25ml
} of 4% (which works perfectly). I am wondering whether my assumption of a
} simple linear relationship between the proportions of NaOH and
} paraformaldehyde is incorrect. Making stock solutions of higher
} concentrations
} does not work either.
}
} Does anyone know a more successful method of preparing large
} volumes of formaldehyde? We do not want to use phosphates or
} preservatives.
}
} regards
} Elizabeth McKenzie
} --------------------------------------------------
} Geomicrobiology and Electron Microscopy Laboratory
} Room S9 Cramer Hall
} 1721 SW Broadway
} Portland State University
} Portland
} OR97201
}
} ph:503 725 3362
} fax:503 725 3025





From: Frank Thomas :      thomasf-at-AGC.BIO.NS.CA
Date: Fri, 23 Jun 2000 07:58:40 -0300
Subject: Strontium-based seashells?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Here's an interesting problem. The other day I was examining some plankton
trap samples in our ESEM, and found a few bits of what I took to be pieces
of foraminiferal shell - they were slightly curved, crystalline-looking
shards with closely spaced fine openings (the foramen, for which forams are
named) and a few large holes (possible spine attachment localities). I ran
EDS on them to make sure that is indeed, what they were. Planktic foram
shells should be fairly pure calcium carbonate. What I got instead were
spectra each showing a fairly large strontium peak, a smaller sulfur peak
and a little oxygen. There were also very high aluminum peaks on each one,
but since the sample was simply air dried on an aluminum stub, I assume
that was the source of the Al.
In 20-some years of micropaleontology I've never heard of any marine
organism that uses strontium to build its shell. Sr is just below Ca in the
periodic table, so I suppose it shares some of the same reactability
characteristics (but I'm no chemist), so perhaps it could substitute, if no
Ca was available (which would be darn funny in a surface marine sample).
Let me add that all the other peaks were in their proper places, and the
instrument had been calibrated quite recently, and I can think of no source
of strontium contamination which could have gotten in there. We're planning
on sending some of these bits out for analysis by some other means just to
make sure, but for the time being, it's quite a mystery.
Any ideas from the marine biologists? The chemists? The EDS gurus?
Psychics?


F.C. Thomas
MicroAnalysis Facility
Geological Survey of Canada (Atlantic)
Bedford Institute of Oceanography
Dartmouth, Nova Scotia
Canada B2Y 4A2




From: Thomas A. Baginski :      tombg-at-bictom.usuf1.usuhs.mil
Date: Fri, 23 Jun 2000 08:09:10 -0400
Subject: ***TEM Darkroom Users*** AGFA Paper has become extinct !

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We use the Agfa DD3700 paper processor. However, the wonderful
graded papers are no longer available and have not been for some time now.
Our back-up system is the Mohr Pro-8. Most of our older trained users are
unwilling to use the AGFA multi-contrast graded papers, stating that the
iamge is "not as crisp, not as white, harder to work with", etc.
If anyone has used AGFA graded papers in the past, and has found a
resin-coated graded paper alternative that works to give sharp, crisp, clean,
white, micrographs from the DURST Point source, I would like to hear about
your solutions.
Some un-tried recommendations have been: FOMA papers, FORTE papers,
ILFORD papers, and then there is the rest of the list, EFKE, Kentmere,
Kodak, MACO, Oriental, adn ECCO. Does anyone know if there is a graded
paper comparable to the AGFA RAPITONE GRADES P1,P2, P 3, and P4. AGFA
only suggests to use their polycontrast resin papers.
I have swithced to digital, but the older folks with all the great
publications refuse to lower their standards.
I would lke to know more about how I can get the wanted details
back onto my micrographs. ANY IDEAS ? ....-TOM

Thomas A Baginski
Technical Coordinator for Microscopy
Uniformed Services University of the Health Sciences
Bethesda, MD 20814-4799

Voice Phone: 301 295 5691
Fax: 301 319 8218
Email: tbaginski-at-usuhs.mil




From: Roger Mason :      rmason-at-sparky2.esd.mun.ca
Date: Fri, 23 Jun 2000 09:55:58 -0230
Subject: Re: Strontium-based seashells?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Frank,

Depending on the polymorph of CaCO3 that foram shells are made of, a few hundred
(calcite) to a few thousand (aragonite) ppm of Sr would not be unusual in carbonate
precipitated from standard sea-water. SO4 ion is also a fairly significant component
of sea-water and S can be incorporated in the structures of carbonate minerals in
small amounts. It could also be present as fluid inclusions. Another possibility is
that the seawater present in the pores of the forams on recovery from the ocean has
precipitated celestite (SrSO4) on dessication, in which case the Sr might have a
very patchy distribution.

Roger Mason

Frank Thomas wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Here's an interesting problem. The other day I was examining some plankton
} trap samples in our ESEM, and found a few bits of what I took to be pieces
} of foraminiferal shell - they were slightly curved, crystalline-looking
} shards with closely spaced fine openings (the foramen, for which forams are
} named) and a few large holes (possible spine attachment localities). I ran
} EDS on them to make sure that is indeed, what they were. Planktic foram
} shells should be fairly pure calcium carbonate. What I got instead were
} spectra each showing a fairly large strontium peak, a smaller sulfur peak
} and a little oxygen. There were also very high aluminum peaks on each one,
} but since the sample was simply air dried on an aluminum stub, I assume
} that was the source of the Al.
}





From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Fri, 23 Jun 2000 08:30:54 -0400
Subject: Re: ***TEM Darkroom Users*** AGFA Paper has become extinct !

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


"Thomas A. Baginski" wrote:

} We use the Agfa DD3700 paper processor. However, the wonderful
} graded papers are no longer available and have not been for some time now.
} Our back-up system is the Mohr Pro-8. Most of our older trained users are
} unwilling to use the AGFA multi-contrast graded papers, stating that the
} iamge is "not as crisp, not as white, harder to work with", etc.
} If anyone has used AGFA graded papers in the past, and has found a
} resin-coated graded paper alternative that works to give sharp, crisp, clean,
} white, micrographs from the DURST Point source, I would like to hear about
} your solutions.
} Some un-tried recommendations have been: FOMA papers, FORTE papers,
} ILFORD papers, and then there is the rest of the list, EFKE, Kentmere,
} Kodak, MACO, Oriental, adn ECCO. Does anyone know if there is a graded
} paper comparable to the AGFA RAPITONE GRADES P1,P2, P 3, and P4. AGFA
} only suggests to use their polycontrast resin papers.
} I have swithced to digital, but the older folks with all the great
} publications refuse to lower their standards.
} I would lke to know more about how I can get the wanted details
} back onto my micrographs. ANY IDEAS ? ....-TOM
}
} Thomas A Baginski
} Technical Coordinator for Microscopy
} Uniformed Services University of the Health Sciences
} Bethesda, MD 20814-4799
}
} Voice Phone: 301 295 5691
} Fax: 301 319 8218
} Email: tbaginski-at-usuhs.mil

Dear Tom:

I have had good results with both Kodak and Ilford multigrade resin coated
papers in my DD-3700. I am surprised that you can see a difference in sharpness
between modern. glossy papers. Is Kodabrom RC still available in grades?

Geoff
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************






From: donald j marshall :      dmrelion-at-world.std.com
Date: Fri, 23 Jun 2000 08:42:39 -0400 (EDT)
Subject: Re: Strontium-bases sea shells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Frank, I am not a microbiologist or even close - I work in
cathodoluminescence instrumentation. I recalled that Sheldon Sommer made
some comments on Sr in shells in his work at Penn State. On page 282 of his
paper, he talks about 70 mole percent strontianite in pelecypods, for what
it is worth. Sommer, S. E. CL of carbonates,2. Geological Applications.
Chemical geology, 1972, pp. 275-284.

Don Marshall


Donald J. Marshall
Relion Industries
P.O. Box 12
Bedford, MA 01730
Ph: 781-275-4695
FAX: 781-271-0252
email dmrelion-at-world.std.com

Cathodoluminescence, mass spectroscopy, electron beam technology


"A weed is a flower out of place."

(Please note: Do not send email with attachments to this address. Instead, send it to donbarlen-at-aol.com. Thank you.)




From: A.P. Alves de Matos :      apmatos-at-ip.pt
Date: Fri, 23 Jun 2000 14:17:25 +0100
Subject: Stainning neurosecretory cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all

I just found some cells with granules resembling neurosecretory granules in
invertebrate tissues, and I need to confirm their nature.

Can anyone recomend a straightforward and reliable method to identify
neurosecretory cells (at the light or/and electron microscopic levels)?


Thanks to all
Dr. A.P. Alves de Matos







From: Philip Oshel :      peoshel-at-facstaff.wisc.edu
Date: Fri, 23 Jun 2000 08:18:36 -0500
Subject: Re: Strontium-based seashells?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There is a group of marine protistans that uses strontium sulphate to
make their tests. Amoeboid critters closely related to ... wait a bit
while I dredge through my memory here, my texts are at home ...
Actinaria? Relatives of the Radiolarians and Heliozoa IF I'm
remembering correctly. They're mentioned inter alia in the book
"Synoptic Classification of Living Organisms" and any good
invertebrate zoology text or Protistology text.

Phil
The Al is most likely from the stub ... try a carbon mount.

} Here's an interesting problem. The other day I was examining
} some plankton
} trap samples in our ESEM, and found a few bits of what I took to be pieces
} of foraminiferal shell - they were slightly curved, crystalline-looking
} shards with closely spaced fine openings (the foramen, for which forams are
} named) and a few large holes (possible spine attachment localities). I ran
} EDS on them to make sure that is indeed, what they were. Planktic foram
} shells should be fairly pure calcium carbonate. What I got instead were
} spectra each showing a fairly large strontium peak, a smaller sulfur peak
} and a little oxygen. There were also very high aluminum peaks on each one,
} but since the sample was simply air dried on an aluminum stub, I assume
} that was the source of the Al.
} In 20-some years of micropaleontology I've never heard of any marine
} organism that uses strontium to build its shell. Sr is just below Ca in the
} periodic table, so I suppose it shares some of the same reactability
} characteristics (but I'm no chemist), so perhaps it could substitute, if no
} Ca was available (which would be darn funny in a surface marine sample).
} Let me add that all the other peaks were in their proper
} places, and the
} instrument had been calibrated quite recently, and I can think of no source
} of strontium contamination which could have gotten in there. We're planning
} on sending some of these bits out for analysis by some other means just to
} make sure, but for the time being, it's quite a mystery.
} Any ideas from the marine biologists? The chemists? The EDS gurus?
} Psychics?
}
}
} F.C. Thomas
} MicroAnalysis Facility
} Geological Survey of Canada (Atlantic)
} Bedford Institute of Oceanography
} Dartmouth, Nova Scotia
} Canada B2Y 4A2

--
}}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
Philip Oshel
Supervisor, AMFSC and BBPIC
Dept. of Animal Health and Biomedical Sciences
University of Wisconsin
1656 Linden Drive
Madison, WI 53706-1581
voice: (608) 263-4162
fax: (608) 262-7420 (dept. fax)




From: barbarac-at-biols.susx.ac.uk ()
Date: Fri, 23 Jun 2000 08:22:55 -0500
Subject: Ask-A-Microscopist:Staining Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form. It was submitted by
(barbarac-at-biols.susx.ac.uk) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday, June 23,
2000 at 08:00:37
---------------------------------------------------------------------------

Email: barbarac-at-biols.susx.ac.uk
Name: Barbara Ciani
School: University of Sussex
Question: Hi everyone,

I need to do some congo red staining on my peptides and
i need to make sure they stick on the microscope slides.

I have been suggested to use gelatine with chrome alum but my slides don't
seem to get 'sticky'.

Any suggestion?

I use 0.5% gelatine (300 bloom) and 0.05% chrome alum.

Thanks,

Barbara

---------------------------------------------------------------------------






From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Fri, 23 Jun 2000 08:39:32 -0500
Subject: ***TEM Darkroom Users*** AGFA Paper has become extinct !

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Date sent: Mon, 19 Jun 2000 09:46:49 -0400
To: {Microscopy-at-sparc5.microscopy.com}
} From: Leona Cohen-Gould {lcgould-at-mail.med.cornell.edu}


Tom,

Although we do very little photographic printing anymore, I printed
extensively in the past and have gotten excellent results from Kodak and
Ilford multi-contrast resin-coated papers, using a Durst professional
enlarger, as well as Beseler enlargers of various types. What I have never
done, though, is make much use of a paper processor. Force of habit, maybe,
but I always preferred tray processing, even when I had to make lots of
prints in short amounts of time.

Even resin-coated papers respond somewhat to variations in developing
times/temperatures, etc., allowing minor adjustments during processing.
Often it is possible to watch the image come up in the tray and know within
15-30 seconds whether it will acceptable or what changes will be necessary.
I can only speak for myself, but quality-wise I believe that tray processing
is superior, in terms of tonality, grey-scale range, etc. I also believe
it's as fast or faster.

That said, I know that people accustomed to processors are just as stubborn
as we dinosaurs who prefer tray processing, and nobody is likely to change
long-established habits. Any mainstream resin-coated, glossy paper should
capture any needed detail that's present in the negative. I'm not saying
that photo paper is as "sharp" in terms of lines/mm as a TEM negative, but
only that the detail it won't capture is too small to see in normal viewing,
and certainly won't survive the reproduction process in a journal. If
you're visibly losing detail in the print at a standard, or even close,
viewing distance without a lens, I would be checking the negatives and
enlarger.

I won't even get into the debate about journals considering digital imaging
as a "lowering of standards"! Not going there... :-)

Just my very subjective two cents.

All the best,
Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/



-----Original Message-----
} From: Thomas A. Baginski [mailto:tombg-at-bictom.usuf1.usuhs.mil]
Sent: Friday, June 23, 2000 7:09 AM
To: Microscopy-at-sparc5.microscopy.com


We use the Agfa DD3700 paper processor. However, the wonderful
graded papers are no longer available and have not been for some time now.

Our back-up system is the Mohr Pro-8. Most of our older trained users are

unwilling to use the AGFA multi-contrast graded papers, stating that the
iamge is "not as crisp, not as white, harder to work with", etc.
If anyone has used AGFA graded papers in the past, and has found a
resin-coated graded paper alternative that works to give sharp, crisp,
clean,
white, micrographs from the DURST Point source, I would like to hear about

your solutions.
Some un-tried recommendations have been: FOMA papers, FORTE papers,

ILFORD papers, and then there is the rest of the list, EFKE, Kentmere,
Kodak, MACO, Oriental, adn ECCO. Does anyone know if there is a graded
paper comparable to the AGFA RAPITONE GRADES P1,P2, P 3, and P4. AGFA
only suggests to use their polycontrast resin papers.
I have swithced to digital, but the older folks with all the great
publications refuse to lower their standards.
I would lke to know more about how I can get the wanted details
back onto my micrographs. ANY IDEAS ? ....-TOM

Thomas A Baginski
Technical Coordinator for Microscopy
Uniformed Services University of the Health Sciences
Bethesda, MD 20814-4799

Voice Phone: 301 295 5691
Fax: 301 319 8218
Email: tbaginski-at-usuhs.mil




From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Fri, 23 Jun 2000 09:46:22 -0400
Subject: Re: Stainning neurosecretory cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


"A.P. Alves de Matos" wrote:

} Dear all
}
} I just found some cells with granules resembling neurosecretory granules in
} invertebrate tissues, and I need to confirm their nature.
}
} Can anyone recomend a straightforward and reliable method to identify
} neurosecretory cells (at the light or/and electron microscopic levels)?
}
} Thanks to all
} Dr. A.P. Alves de Matos

Chrome-alum hematoxylin or Falhmi's aldehyde fuchsin work well with mammalian
tissues. Check almost any edition of Humason's "Animal Tissue Techniques" for
specifics. Modifications for invert tissues are out there but I don't have
references handy.

Geoff
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************






From: Mike Coviello :      coviello-at-mae.uta.edu
Date: Fri, 23 Jun 2000 09:23:02 -0500
Subject: Re: TEM-Indepedent FIB lab-a thank you

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Y'all:
Thanks for your overwelming response to my request for an independent
FIB lab that can do TEM x-sections. Over the years, this forum has been
an invaluable resource for me.
Regards, Mike Coviello
Lab Manager
University of Texas -at- Arlington





From: Belluso elena :      belluso-at-dsmp.unito.it
Date: Fri, 23 Jun 2000 16:44:22 +0200 (MDT)
Subject: Need mineralogical samples for TEM studies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everyone,
I'm studying microstructures and intergrowths of fibrous minerals by TEM.
I'm looking for a sample of amosite and a sample of antophillite fibres in a
rock matrix in order to continue this study. I'm gratefully if someone will
send my these samples and will be interested to collaborate.

Thank you everyone,
Elena Belluso



----------------------------------------------------
Elena BELLUSO
Dipartimento di Scienze Mineralogiche e Petrologiche
Via Valperga Caluso, 35
I-10125 TORINO - ITALIA
tel:(39) 011.670.71.35 - fax: (39) 011.670.71.28
e-mail: belluso-at-dsmp.unito.it
http://www.dsmp.unito.it
----------------------------------------------------





From: tracy gales :      tl_gales-at-fccc.edu
Date: Fri, 23 Jun 2000 11:01:58 -0400
Subject: Please unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




--
Tracy Gales
Fox Chase Cancer Center
Electron Microscopy Facility
7701 Burholme Ave.
Philadelphia PA 19111
Tel. 215-728-2484
Fax 215-728-2412






From: Stuart McKernan :      mckernan-at-cems.umn.edu
Date: Fri, 23 Jun 2000 10:36:55 -0500 (CDT)
Subject: M&M 2000; Late-breaking poster session

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The deadline for contributions to the "LATE-BREAKING POSTER SESSION" at
MICROSCOPY AND MICROANALYSIS 2000 is very close at hand.

This poster session will be composed of presentations of newly acquired data or
analyses which were unavailable for submission by the February 15 deadline. A
short, half page abstract describing the studies is required. The abstract
should include: Title, Authors, Authors affiliation, and a Brief Description of
the studies. The description should include the Aim of the studies, a short
characterization of the Methods, and a brief account of the Results and their
Importance. The abstracts will obviously not be published in the proceedings,
since it is already being printed, but will be available at the meeting. The
posters will be advertized in the daily meeting newsletters to alert attendees
to their presence.

Abstracts should be e-mailed or faxed to the program chair, Stuart McKernan, at
stuartm-at-tc.umn.edu (email) or 612-625-5368 (fax). Abstracts must be received by
June 23, 2000. Abstracts will be reviewed by members of the program committee. A
limited number of poster boards are available and preference will be given to
early submissions. Abstract authors will be notified of acceptance of their
abstracts no later than July 1 (earlier for early submissions).

It is not too early to register for the meeting, the pre-meeting congress, or
the short courses (or any combination of these events). The early registration
deadline is also fast approaching, and can be done using the online registration
site at: http://www.peregrine.net/mm2000/

Looking forward to a great meeting in Philadelphia,



__________________
Stuart McKernan stuartm-at-tc.umn.edu
Director Office:(612) 626-7594
IT Characterization Facility, University of Minnesota Desk:(612) 624-6009
100 Union Street S. E., Minneapolis, MN 55455 NEW-} Fax:(612) 625-5368





From: Belluso elena :      belluso-at-dsmp.unito.it
Date: Fri, 23 Jun 2000 18:13:02 +0200 (MDT)
Subject: Need mineralogical samples for TEM studies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everyone,
I'm studying microstructures and intergrowths of fibrous minerals by TEM.
I'm looking for a sample of amosite and a sample of antophillite fibres in a
rock matrix in order to continue this study. I'm gratefully if someone will
send my these samples and will be interested to collaborate.

Thank you everyone,
Elena Belluso



----------------------------------------------------
Elena BELLUSO
Dipartimento di Scienze Mineralogiche e Petrologiche
Via Valperga Caluso, 35
I-10125 TORINO - ITALIA
tel:(39) 011.670.71.35 - fax: (39) 011.670.71.28
e-mail: belluso-at-dsmp.unito.it
http://www.dsmp.unito.it
----------------------------------------------------





From: Matthew Lynn :      mlynn-at-miami.edu
Date: Fri, 23 Jun 2000 12:26:59 -0400
Subject: RE: Strontium-based seashells?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Frank,

Yes...very interesting result! My guess is that you have happened upon a very
rare group of microplankton which make a celestite test (SrSO4). If I remember
my micropaleo, look for a group called Acantharia (?) Try Haq & Boersma's
"Marine Micropaleontology" for a start. The reason you have not come across
them in 20-odd years is that the celestite tests are very unstable...I saw a
few beautiful examples in filtered water from the Arabian Sea, never in
sediment samples. So not much use for paleoclimatology, but an interesting
diversion. If you can post some pics to a website, I'd like to take a look at
them.

Matt

Matthew J. Lynn, Ph.D.
Center for Advanced Microscopy
University of Miami
(305)284-4736
mlynn-at-miami.edu


On Friday, June 23, 2000 6:59 AM, Frank Thomas [SMTP:thomasf-at-AGC.BIO.NS.CA]
wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Here's an interesting problem. The other day I was examining some plankton
} trap samples in our ESEM, and found a few bits of what I took to be pieces
} of foraminiferal shell - they were slightly curved, crystalline-looking
} shards with closely spaced fine openings (the foramen, for which forams are
} named) and a few large holes (possible spine attachment localities). I ran
} EDS on them to make sure that is indeed, what they were. Planktic foram
} shells should be fairly pure calcium carbonate. What I got instead were
} spectra each showing a fairly large strontium peak, a smaller sulfur peak
} and a little oxygen. There were also very high aluminum peaks on each one,
} but since the sample was simply air dried on an aluminum stub, I assume
} that was the source of the Al.
} In 20-some years of micropaleontology I've never heard of any marine
} organism that uses strontium to build its shell. Sr is just below Ca in the
} periodic table, so I suppose it shares some of the same reactability
} characteristics (but I'm no chemist), so perhaps it could substitute, if no
} Ca was available (which would be darn funny in a surface marine sample).
} Let me add that all the other peaks were in their proper places, and the
} instrument had been calibrated quite recently, and I can think of no source
} of strontium contamination which could have gotten in there. We're planning
} on sending some of these bits out for analysis by some other means just to
} make sure, but for the time being, it's quite a mystery.
} Any ideas from the marine biologists? The chemists? The EDS gurus?
} Psychics?
}
}
} F.C. Thomas
} MicroAnalysis Facility
} Geological Survey of Canada (Atlantic)
} Bedford Institute of Oceanography
} Dartmouth, Nova Scotia
} Canada B2Y 4A2




From: Anthony Garratt-Reed :      tonygr-at-mit.edu
Date: Fri, 23 Jun 2000 11:57:45 -0400
Subject: Re: Hole in X-ray spectra

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Oldrich,

I think you have massive icing of your detector crystal, or equally massive
contamination of your window. I have modelled your Al spectrum (using
DTSA) and can get quite good agreement with your spectrum by putting in 300
microns of ice in the model. The low energy rise is probably caused simply
by the tail of the electronic noise which is amplified so much by the scale
at which you display your spectrum. Because of the way x-ray absorption
works, I can't really tell the difference between absorption in ice and
absorption in oil. Anyway, fixing an icing problem is far easier than
fixing an oil problem - you simply have to warm up the detector while
pumping on it. This will need an adaptor to fit on the pumping port of
your detector - your service engineer should have, or be able to get hold
of, one of these. So I would try warming the detector and seeing of it
solved the problem.

I wouldn't stake my professional reputation on this diagnosis, but I am
certain that if the problem is not contamination, then it is a geometrical
problem - the collimator may have moved, or some other mechanical
derangement taken place.

Good luck,

Tony Garratt-Reed.


At 11:11 AM 06/23/2000 +0200, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6479
**






From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Fri, 23 Jun 2000 09:48:06 -0700
Subject: Re: undergrad SEM use

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ken,
I have always run undergraduate courses on the SEM/EDX and we currently run
a course that puts groups of second year engineering students doing
independent research projects. They research the composition of many common
household and commercial appliances, taking them apart to see what metals,
plastics and ceramics are used for the various applications such as small
electric motors, heating elements and temperature control. The students are
very appreciative of being allowed to sit down and use the instrument with
minimal supervision and I find they are very careful with it.
I run courses for Anthropology and Electrical Engineering and all
disciplines of Engineering, Physics and Chemistry use the instrument for
both undergraduate and graduate research. I would suggest you get an EDX and
a low vacuum ("environmental") SEM for full versatility in both physical and
biological EM.
All of my EM instruments (two SEMs with EDX, one 200 kV TEM and SEM with
EBSP) are Hitachi, as I have found them the most robust. They run for years
of student use and nothing ever seems to go wrong with them. I had an ETEC
before my first Hitachi SEM and I liked it too, but they have been out of
business for a long time.
We also have a fourth year course that covers XRD, SEM, EDX and TEM/EDX. The
professor in charge has developed the lab course to touch all of the
important principles and demonstrate them in the lab. We cover changing kV,
EDX excitation volume, EDX matrix corrections, TEM use and TEM/EDX issues.
Please contact me if I can give you any other information.
At 03:05 PM 6/22/00 -0700, you wrote:
}
} Greetings:
} I am a new subscriber to this list and am looking for input/advice
} regarding the use of SEM's in undergraduate research. I teach at a liberal
} arts university with only undergraduate programs in the natural sciences.
} A geology colleague and myself (I'm a biologist) are exploring possible
} funding sources which would allow us to purchase an SEM that would be used
} by biology and geology undergrads in their independent and senior thesis
} projects. (We currently have a non-functioning ETEC that we would like to
} replace -- we've been advised that further repair and head-banging would
} not be cost effective). If anyone is currently engaged in SEM-based
} research with undergrads, I would like to get your feedback on your
} experiences (both the pluses and minuses). Although we don't have a TEM,
} input from those using TEM with undergrads would also be helpful.
}
} Specific Questions:
}
} 1) What types of projects are your students pursuing?
}
} 2) Are you conducting interdisciplinary projects? e.g., micropaleontology
} or environmental applications.
}
} 3) Are there particular models of SEMs which would be more user friendly
} since we would be training undergrads and we would also be the primary
} trainers and technicians?
}
} 4) Have you developed an undergrad EM course?
}
} Thanks,
} Ken

Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca





From: Robert Santoianni :      Robert_Santoianni-at-emory.org
Date: Fri, 23 Jun 2000 13:00:49 -0500
Subject: Re:TEM Darkroom/AGFA Papers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Date sent: Fri, 23 Jun 2000 11:57:42 -0400 (EDT)
} From: Donald Lovett {lovett-at-TCNJ.EDU}
To: Chris Jeffree {cjeffree-at-srv0.bio.ed.ac.uk}


When AGFA (Bayer) pulled the plug on the graded papers, we switched to
Kodak Polycontrast RCIII. So I pulled the plug on their activator and
fixer and now use Kodak Polymax developer 1:4 dilution and Rapid Fixer
in my DD3700. Still use my Durst S45EM with point light source and
Kodak Polycontrast Filter Kit.
Bob Santoianni
Emory University Hospital
Atlanta, GA




From: Caroline Schooley :      schooley-at-mcn.org
Date: Fri, 23 Jun 2000 10:14:58 -0700
Subject: Re: Strontium-based seashells?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Here's an interesting problem. The other day I was examining some
} plankton
} trap samples in our ESEM, and found a few bits of what I took to be pieces
} of foraminiferal shell - they were slightly curved, crystalline-looking
} shards with closely spaced fine openings (the foramen, for which forams are
} named) and a few large holes (possible spine attachment localities). I ran
} EDS on them to make sure that is indeed, what they were. Planktic foram
} shells should be fairly pure calcium carbonate. What I got instead were
} spectra each showing a fairly large strontium peak, a smaller sulfur peak
} and a little oxygen. There were also very high aluminum peaks on each one,
} but since the sample was simply air dried on an aluminum stub, I assume
} that was the source of the Al.
} In 20-some years of micropaleontology I've never heard of any marine
} organism that uses strontium to build its shell. Sr is just below Ca in the
} periodic table, so I suppose it shares some of the same reactability
} characteristics (but I'm no chemist), so perhaps it could substitute, if no
} Ca was available (which would be darn funny in a surface marine sample).
} Let me add that all the other peaks were in their proper places,
} and the
} instrument had been calibrated quite recently, and I can think of no source
} of strontium contamination which could have gotten in there. We're planning
} on sending some of these bits out for analysis by some other means just to
} make sure, but for the time being, it's quite a mystery.
} Any ideas from the marine biologists? The chemists? The EDS gurus?
} Psychics?

Frank -

You've just rediscovered the strontianate Radiolaria. They DO selectively
concentrate strontium. I almost selected them as my thesis topic 'way
back in the dark ages; I dropped the idea when I realised that there was no
way to culture them in the lab. I don't think there is, yet.
}
Caroline

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: Doug Cromey :      doug-cromey-at-ns.arizona.edu
Date: Fri, 23 Jun 2000 10:36:41 -0700
Subject: Re: Stainning neurosecretory cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This may be a bit of a shameless plug for a paper I contributed to, but
here's the reference:

Payne CM, Cromey DW (1987) Ultrastructural similarities between Chlorohydra
viridissima and human neurosecretory granules: A cytochemical study using
the uranaffin reaction. Cytobios 50:191-203.

Back when I was doing clinical TEM, we used the Uranaffin Method as the
"gold standard" for diagnostic identification of neurosecretory granules in
tumors. It was more reliable than immunomarkers and its fairly simple to
do. The reference section from this particular article should include Dr.
Payne's original publication on the technique. The most important thing is
to follow the protocol, failure to use the specified buffers usually causes
problems.

Yours,
Doug Cromey

At 09:46 AM 6/23/2000 -0400, you wrote:
} "A.P. Alves de Matos" wrote:
} } I just found some cells with granules resembling neurosecretory granules in
} } invertebrate tissues, and I need to confirm their nature.
} }
} } Can anyone recomend a straightforward and reliable method to identify
} } neurosecretory cells (at the light or/and electron microscopic levels)?
}
} Chrome-alum hematoxylin or Falhmi's aldehyde fuchsin work well with mammalian
} tissues. Check almost any edition of Humason's "Animal Tissue Techniques" for
} specifics. Modifications for invert tissues are out there but I don't have
} references handy.

....................................................................
: Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy :
: Research Specialist, Principal University of Arizona :
: (office: AHSC 4212A) P.O. Box 245044 :
: (voice: 520-626-2824) Tucson, AZ 85724-5044 USA :
: (FAX: 520-626-2097) (email: doug-cromey-at-ns.arizona.edu):
:...................................................................:
http://www.pharmacy.arizona.edu/exp_path.html
Home of: "Microscopy and Imaging Resources on the WWW"





From: Matthew Lynn :      mlynn-at-miami.edu
Date: Fri, 23 Jun 2000 13:37:46 -0400
Subject: Re: Strontium-based seashells?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry if this is double-posted....server troubles!

Frank,

Yes...very interesting result! My guess is that you have happened upon a very
rare group of microplankton which make a celestite test (SrSO4). If I remember
my micropaleo, look for a group called Acantharia (?) Try Haq & Boersma's
"Marine Micropaleontology" for a start. The reason you have not come across
them in 20-odd years is that the celestite tests are very unstable...I saw a
few beautiful examples in filtered water from the Arabian Sea, never in
sediment samples. So not much use for paleoclimatology, but an interesting
diversion. If you can post some pics to a website, I'd like to take a look at
them.

Matt


On Friday, June 23, 2000 6:59 AM, Frank Thomas [SMTP:thomasf-at-AGC.BIO.NS.CA]
wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Here's an interesting problem. The other day I was examining some plankton
} trap samples in our ESEM, and found a few bits of what I took to be pieces
} of foraminiferal shell - they were slightly curved, crystalline-looking
} shards with closely spaced fine openings (the foramen, for which forams are
} named) and a few large holes (possible spine attachment localities). I ran
} EDS on them to make sure that is indeed, what they were. Planktic foram
} shells should be fairly pure calcium carbonate. What I got instead were
} spectra each showing a fairly large strontium peak, a smaller sulfur peak
} and a little oxygen. There were also very high aluminum peaks on each one,
} but since the sample was simply air dried on an aluminum stub, I assume
} that was the source of the Al.
} In 20-some years of micropaleontology I've never heard of any marine
} organism that uses strontium to build its shell. Sr is just below Ca in the
} periodic table, so I suppose it shares some of the same reactability
} characteristics (but I'm no chemist), so perhaps it could substitute, if no
} Ca was available (which would be darn funny in a surface marine sample).
} Let me add that all the other peaks were in their proper places, and the
} instrument had been calibrated quite recently, and I can think of no source
} of strontium contamination which could have gotten in there. We're planning
} on sending some of these bits out for analysis by some other means just to
} make sure, but for the time being, it's quite a mystery.
} Any ideas from the marine biologists? The chemists? The EDS gurus?
} Psychics?
}

Matthew J. Lynn, Ph.D.
Center for Advanced Microscopy
University of Miami
(305)284-4736
mlynn-at-miami.edu





From: Matt Ervin :      mervin-at-ARL.mil
Date: Fri, 23 Jun 2000 13:38:27 -0400
Subject: A hole in the EDX spectrum II

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Oldrich-
I am sorry, but I have forgotten all of the details of your problem.
Have you thermally cycled your detector? letting the detector be a room
temperature for at least 24hrs can remove ice build-up in the detector. It
can also be useful to purge the dewar with nitrogen at that time to remove
any ice that may be in the dewar thus reducing cooling efficiency. I would
also suggest cleaning your detector window (I don't know what kind you
have.) Be very sure that you are using the proper technique and chemicals
for cleaning the window as the polymer windows are extremely easy to damage
(i.e., don't blow a dust particle off of it, see the current thread on
particulate damage to EDX windows.) In my case, as my window gets coated
with pump oil from the vacuum, etc. the Cu L line reduces in intensity
relative to the K line. For my super thin window the ratio should be L:K
} =2:1. If I put a thin graphite sheet between the detector and the sample,
I get spectra very similar to the ones you have posted with the low energy
x-rays blocked out (you'll still have the electronic noise peak at the
lowest energy) and then a normal looking higher energy range. Perhaps some
of this will help. I'll be on vacation the next couple of weeks, but give
me a call after 7 July if you have any questions I can help with.
Matthew Ervin
MErvin-at-ARL.mil
301-394-0017




Oldrich Benada {benada-at-biomed.cas.cz} on 06/23/2000 05:11:15 AM





To: Microscopy-at-sparc5.microscopy.com
cc:


Dear colleagues,
Many thanks for all responses to my question considering the hole in
the EDX spectrum.
We had a service engineer in our lab and he tried to solve the
problem but he was not successful.
I have prepared WEB page with the images of spectra, that were
recorded from following samples:
Cu_Al sample (usually used for calibration)
Ti grid
AL grid
K standard

The url of that page is following:
http://www.biomed.cas.cz/~benada/EDX_page/index.htm

Please, if anyone would be so kind and could comment the spectra, we
will be very happy.
Best regards from Prague
Oldrich Benada


+-----------------------------------+
Oldrich Benada
Acad. Sci. CR
Institute of Microbiology
Laboratory of electron microscopy
Videnska 1083
CZ - 142 20 Prague 4 - Krc
Czech Republic
+------------------------------------+
Phone: +420-2-4752399
Fax: +420-2-4715743
WEB:http://www.biomed.cas.cz/mbu/lem113/lem.htm








From: sherwood-at-helix.mgh.harvard.edu (Peggy Sherwood)
Date: Fri, 23 Jun 2000 13:45:54 -0400
Subject: AGFA Paper alternatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To all:
We have been using Kodabrome II RC paper F1-F4 for years with our Durst
Laborator 1200 and
Rapidoprint DD3700. As of the end of 1999, the paper is still available
and gives good results.

Peggy Sherwood
Photopathology
Wellman Labs of Photomedicine-W224
70 Blossom Street
Boston, MA 02114

617-726-6983 (Lab)
617-724-4839 (Voice)
617-726-3192 (Fax)






From: Brian McIntyre :      mcintyre-at-optics.rochester.edu
Date: Fri, 23 Jun 2000 14:14:24 -0400
Subject: coin question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all-

Does anyone know how a really good counterfeiter would make a change in the
date field of a rare coin look as though it weren't altered? I've looked
for markings associated with moving the metal around, a joint between the
numbers and the base coin, and differing metallurgy, with no "smoking gun"....

Any thoughts??

Brian
----------------------------------------------
Brian McIntyre
Electron Microscopy Lab- River Campus
Univ of Rochester
Rochester, NY 14627
716-275-3058/4875




From: Matthew Lynn :      mlynn-at-miami.edu
Date: Fri, 23 Jun 2000 14:06:44 -0400
Subject: Re: Strontium-based seashells?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sorry if this is double-posted....server troubles!

Frank,

Yes...very interesting result! My guess is that you have happened upon a very
rare group of microplankton which make a celestite test (SrSO4). If I remember
my micropaleo, look for a group called Acantharia (?) Try Haq & Boersma's
"Marine Micropaleontology" for a start. The reason you have not come across
them in 20-odd years is that the celestite tests are very unstable...I saw a
few beautiful examples in filtered water from the Arabian Sea, never in
sediment samples. So not much use for paleoclimatology, but an interesting
diversion. If you can post some pics to a website, I'd like to take a look at
them.

Matt


On Friday, June 23, 2000 6:59 AM, Frank Thomas [SMTP:thomasf-at-AGC.BIO.NS.CA]
wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Here's an interesting problem. The other day I was examining some plankton
} trap samples in our ESEM, and found a few bits of what I took to be pieces
} of foraminiferal shell - they were slightly curved, crystalline-looking
} shards with closely spaced fine openings (the foramen, for which forams are
} named) and a few large holes (possible spine attachment localities). I ran
} EDS on them to make sure that is indeed, what they were. Planktic foram
} shells should be fairly pure calcium carbonate. What I got instead were
} spectra each showing a fairly large strontium peak, a smaller sulfur peak
} and a little oxygen. There were also very high aluminum peaks on each one,
} but since the sample was simply air dried on an aluminum stub, I assume
} that was the source of the Al.
} In 20-some years of micropaleontology I've never heard of any marine
} organism that uses strontium to build its shell. Sr is just below Ca in the
} periodic table, so I suppose it shares some of the same reactability
} characteristics (but I'm no chemist), so perhaps it could substitute, if no
} Ca was available (which would be darn funny in a surface marine sample).
} Let me add that all the other peaks were in their proper places, and the
} instrument had been calibrated quite recently, and I can think of no source
} of strontium contamination which could have gotten in there. We're planning
} on sending some of these bits out for analysis by some other means just to
} make sure, but for the time being, it's quite a mystery.
} Any ideas from the marine biologists? The chemists? The EDS gurus?
} Psychics?
}

Matthew J. Lynn, Ph.D.
Center for Advanced Microscopy
University of Miami
(305)284-4736
mlynn-at-miami.edu





From: Earl Weltmer :      earlw-at-pacbell.net
Date: Fri, 23 Jun 2000 18:24:16 -0700
Subject: Re: coin question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Brian,

I've thought about this before.

A really good counterfeiter could do this at the molecular level by using and FIB
or equivalent.

I haven't really explored exactly how this could be done but it has crossed my
mind.


regards,

Earl Weltmer

Brian McIntyre wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hi all-
}
} Does anyone know how a really good counterfeiter would make a change in the
} date field of a rare coin look as though it weren't altered? I've looked
} for markings associated with moving the metal around, a joint between the
} numbers and the base coin, and differing metallurgy, with no "smoking gun"....
}
} Any thoughts??
}
} Brian
} ----------------------------------------------
} Brian McIntyre
} Electron Microscopy Lab- River Campus
} Univ of Rochester
} Rochester, NY 14627
} 716-275-3058/4875





From: Rajesh Patel :      rpatel-at-UMDNJ.EDU
Date: Sat, 24 Jun 2000 12:39:21 -0500
Subject: stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There was a recent posting about UA precipitating during staining.

Well wouldn't you know I too have experinecing similar problem. It
took me a long time to figure out but I narrowed it down to our
rinsing water that is used to rinse the grid in after it is stained.

After chaging the filters on our water purification system the problem was
gone.

Rajesh Patel
Robert Wood Johnson Medical School
Dept. of Pathology
Electron Microscopy Lab
675 Hoes Lane
Piscataway, NJ 08854

(732)235-4648
rpatel-at-umdnj.edu






From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Sat, 24 Jun 2000 14:14:10 -0600
Subject: coin question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Brian,

how about looking at sub-surface damage (using some channeling techniques or
similar)? I could imagine, that the damage from stamping the coin is
different from Ion bombarding the coin.

Or, look for implanted atoms. The FIB process uses metal atoms to bombard
the sample. Some of them get implanted into the material. So you might find
higher concentrations where the sample was altered.

Finally, how about a surface layer like an oxide or patina. There may be
differences between the very old layer and the new layer.

Michael

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Brian McIntyre [mailto:mcintyre-at-optics.rochester.edu]
Sent: Friday, June 23, 2000 12:14 PM
To: Microscopy-at-sparc5.microscopy.com


Hi all-

Does anyone know how a really good counterfeiter would make a change in the
date field of a rare coin look as though it weren't altered? I've looked
for markings associated with moving the metal around, a joint between the
numbers and the base coin, and differing metallurgy, with no "smoking
gun"....

Any thoughts??

Brian
----------------------------------------------
Brian McIntyre
Electron Microscopy Lab- River Campus
Univ of Rochester
Rochester, NY 14627
716-275-3058/4875




From: Jo Verbeeck :      joverbee-at-ruca.ua.ac.be
Date: Mon, 26 Jun 2000 09:29:37 +0200 (METDST)
Subject: Power stability in microscopy and EELS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

Over time I saw some questions regarding the installation of a UPS or some
power conditioner to improve the stability of all sorts of microscopy
equipment.
I understand that quite a few instruments are sensitive to the quality of
the line power they get.
As I was working on EMC (electro magnetic compatibility) in a former life,
this seems strange to me.
In Europe we have very strict EC-rules that guarantee that equipment meets
its specifications even under severe conditions (including HF noise on the
power, fluctuations in the power, high EM fields, temperature variations
etc)

I agree that microscopes are far more sensitive devices than, say,
videogames but does anyone know of the EMC rules for microscopes and
related equipment?
It seems strange that some people are using UPS's while especially power
fluctuations can be easily overcome with a good design of the
electronics...

Regards,

Jo

*************************************************************
* Jo Verbeeck *
* University of Antwerp *
* Dept. EMAT (Electron Microscopy for Materials Research) *
* e-mail: joverbee-at-ruca.ua.ac.be *
* tel: +32(0)3 218 02 49 *
* fax: +32(0)3 218 02 57 *
*************************************************************





From: oshel-at-terracom.net (Philip Oshel)
Date: Mon, 26 Jun 2000 06:01:20 -0500
Subject: protists & SrSO4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I've got my references now:
The protists with SrSO4 tests are the class *Acantherea*, members of the
phylum Sarcodina (amoebae), subphylum Actinopoda (used to be known as the
Radiolaria).

By my somewhat outdated reference ("A Synoptic Classification of Living
Organisms", R.S.K. Barnes, ed.). The names have likely been changed, and
the Actinopoda broken up to protect the careers of systematists.

Forget I ever mentioned Actinaria or Acanthella. The long-term storage in
my wetware is getting flakey (I misremembered).

Phil

}}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
Philip Oshel
Supervisor, AMFSC and BBPIC
Dept. of Animal Health and Biomedical Sciences
University of Wisconsin
1656 Linden Drive
Madison, WI 53706-1581
Voice: (608) 263-4162
peoshel-at-facstaff.wisc.edu
fax: (608) 262-7420 (dept. fax)






From: HARRISm-at-esm-semi.co.uk
Date: Mon, 26 Jun 2000 12:04 +0000 (GMT)
Subject: Electron Microscope Maintenance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi ,


Could anyone recommend an independent U.K based company
specialising in the repair and maintenance of electron microscopes ?

Without mentioning names I feel that the service we receive re our SEM
maintenance contract is particularly poor and am beginning to wonder
is there any alternative to complaint without improvement and excuse
after excuse .

I have noticed that little repair is carried out at component level
during service engineer visits but am not aware of how ' specific '
most parts are to a particular make of instrument . If parts always
have to be ordered through the instrument supplier there would be no
advantage in independent maintenance .
Another alternative perhaps would be in-house maintenance but do
courses exist to train the novice type like myself to a reasonably
competent level or is this not advised ?

Keen to hear comments ...

M.Harris Email harrism-at-esm-semi.co.uk
ESM LTD ,
South Wales , U.K







From: Rudi Lurz :      Lurz-at-molgen.mpg.de
Date: Mon, 26 Jun 2000 13:02:29 +0200
Subject: RE: AGFA Paper

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Date: Mon, 26 Jun 2000 09:56:43 +0200
} To: Microscopy-at-sparc5.microscopy.com
} From: Rudi Lurz {Lurz-at-molgen.mpg.de}
} Subject: Re: ***TEM Darkroom Users*** AGFA Paper has become extinct !
}
} To all:
} The identical technology (developer in emulsion + activator) as with
} Rapidoprint papers is used with Agfa Brovira-Speed papers. I have replaced
} years ago the Rapitone paper by Brovira-Speed 310 RC because I prefered
} more the blue-black of Brovira-Speed to Rapitone which showed more brown
} tendency. Brovira-Speed is available from soft to extra hard and should be
} still produced (hopefully) as I was told from my dealer this morning.
} This is for Germany - I do not know the situation in the US.
} Agfa seems to reduce drastically the B/W programme. The Agfa Scientia
} 23D56 films are no longer produced too and we have to switch to Kodak SO-163.
} Best regards
} Rudi Lurz


_________________________________
Rudi Lurz Phone: X - 30-8413-1271
MPI fźr Molekulare Genetik Fax: X - 30-8413-1385
Ihnestrasse 73
D-14195 Berlin E-mail to: Lurz-at-molgen.mpg.de







From: Philip Oshel :      peoshel-at-facstaff.wisc.edu
Date: Mon, 26 Jun 2000 08:15:18 -0500
Subject: Re: undergrad SEM use

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Teaching microscopy at both the undergrad and grad student levels
would be an excellent symposium or discussion group at the Long Beach
meeting. Especially since the life of EM facilities depends on having
users, and getting students interested in microscopy is one of the
best ways to create users.

Phil

} } Greetings:
} } I am a new subscriber to this list and am looking for input/advice
} } regarding the use of SEM's in undergraduate research. I teach at a liberal
} } arts university with only undergraduate programs in the natural sciences.
} } A geology colleague and myself (I'm a biologist) are exploring possible
} } funding sources which would allow us to purchase an SEM that would be used
} } by biology and geology undergrads in their independent and senior thesis
} } projects. (We currently have a non-functioning ETEC that we would like to
} } replace -- we've been advised that further repair and head-banging would
} } not be cost effective). If anyone is currently engaged in SEM-based
} } research with undergrads, I would like to get your feedback on your
} } experiences (both the pluses and minuses). Although we don't have a TEM,
} } input from those using TEM with undergrads would also be helpful.
} }
} } Specific Questions:
} }
} } 1) What types of projects are your students pursuing?
} }
} } 2) Are you conducting interdisciplinary projects? e.g., micropaleontology
} } or environmental applications.
} }
} } 3) Are there particular models of SEMs which would be more user friendly
} } since we would be training undergrads and we would also be the primary
} } trainers and technicians?
} }
} } 4) Have you developed an undergrad EM course?
} }
} } Thanks,
} } Ken
--
}}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
Philip Oshel
Supervisor, AMFSC and BBPIC
Dept. of Animal Health and Biomedical Sciences
University of Wisconsin
1656 Linden Drive
Madison, WI 53706-1581
voice: (608) 263-4162
fax: (608) 262-7420 (dept. fax)




From: Brian McIntyre :      mcintyre-at-optics.rochester.edu
Date: Mon, 26 Jun 2000 11:22:39 -0400
Subject: coin question followup

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


hi...and thanks for the helpful insights thusfar.

i should clarify that the coin in question could not have been altered
after 1900 so modern techniques are obviously out of the question. the
issue becomes one of... if you wanted to alter a coin's date how would you
have done it 100 years ago, and how could it be uncovered using modern
analysis techniques.

thanks!
b-
----------------------------------------------
Brian McIntyre
Electron Microscopy Lab- River Campus
Univ of Rochester
Rochester, NY 14627
716-275-3058/4875




From: Allen R. Sampson :      ars-at-sem.com
Date: Mon, 26 Jun 2000 11:33:13 -0500
Subject: RE: coin question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Use of an x-ray microprobe should be rather definitive, I would think.
While there are exotic (and expensive) techniques that could be used to
remove a raised date and place a new one, it would be extremely difficult
to exactly match the elemental contents, particlarly those of the trace
elements. For example, gold, even when refined to 99.999% pure, contains
enough impurities for any batch to be matched to a particular mine that it
came from.

The only other method I could imagine is a close characterization of the
grain structures. I don't know if the stamping process would have much of
an affect on the surface grain, but I imagine any implantation method would
leave a much different structure as completely different conditions were
used to form the metal.



Allen R. Sampson, Owner
Advanced Research Systems, St. Charles, Illinois 60174
voice 630.513.7093 fax 630.513.7092

On Friday, June 23, 2000 1:14 PM, Brian McIntyre
[SMTP:mcintyre-at-optics.rochester.edu] wrote:

} Hi all-
}
} Does anyone know how a really good counterfeiter would make a change in
the
} date field of a rare coin look as though it weren't altered? I've looked
} for markings associated with moving the metal around, a joint between the
} numbers and the base coin, and differing metallurgy, with no "smoking
gun"....
}
} Any thoughts??
}
} Brian
} ----------------------------------------------
} Brian McIntyre
} Electron Microscopy Lab- River Campus
} Univ of Rochester
} Rochester, NY 14627
} 716-275-3058/4875
}
}




From: Mark West :      mwest-at-ifcsun1.ifisiol.unam.mx
Date: Mon, 26 Jun 2000 13:36:33 -0600 (Hora est‡ndar de MŽxico)
Subject: frogs' eggs and vesicles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

A couple of weeks ago I wrote in witha problem trying to fix plant vesicle
preps for LR White inclusion and immunogold studies, without osmicating.
The pellets were disappearing during dehydration. I included an osmium
step (5, 30 or 60 minutes) and the pellets stopped disappearing, so I
guess there is so little protein that a formaldehyde/gluteraldehyde fix
doesn't manage to stabilize them. I'll have to check whether these
osmicated vesicles work in the immunogold reaction and check the effect
of a metaperiodate treatment to expose the proteins. Thanks for the advice
- another problem solved.

Someone brought us some frogs' eggs to fix. It looks like they have a hard
skin. The researcher is interested in extracting the nuclei (huge) which
also look like they're enclosed in a hard skin, and he wants to know
whether the extracted nuclei still retain some endoplasmic reticulum, so I
thought the best way would be to compare a whole egg with an extracted
nucleus. Any ideas/experience on how to fix this thing?

Thanks,

Mark






********************************************
Mark West,
Unidad de Microscopia Electronica,
(Electron Microscopy Unit)
Instituto de Fisiologia Celular,
Universidad Nacional Autonoma de Mexico,
04510 Mexico D.F.

tel (unidad/lab) *(525) 622 5610*
(casa/home) (525) 619 3020
Fax (525) 616 2282
********************************************





From: Heejun Jeong :      hjjeong -at-physics.purdue.edu
Date: Tue, 27 Oct 1998 14:47:09 -0500
Subject: question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I just want some opinions about the problem I got recently with our SEM.

We use this machine for electron beam lithography. This application is a
little more sensitive than imaging. The problem I have now is, when I
image at very low magnification, I can see the circular rim of the final
section of the column. But from a couple of weeks, it changed suddenly,
now the circular image is distorted like an oval shape and at second
saturation point, normally only one position of gun alignment gives the
maximam beam current, but I see there are clearly two distinct, well
separated positions. We think something is blocking in the middle of the
beam path, but want to know whether there is any other one who
experienced similar thing and know the reason.
I appreciate if anyone can give me his/her similar experince.

Heejun Jeong
hjjeong-at-physics.purdue.edu






From: Orion Analytical :      orionanalytical-at-worldnet.att.net
Date: Mon, 26 Jun 2000 15:47:14 -0400
Subject: Re: coin question followup

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Coins have been faked outright and updated using modern dies, and assembled
from two separate halves using solder. Coins whose value depends on date
have been "enriched" by removing the original date and soldering in its
place numbers of a more valuable date. This latter case might be discerned
using scanning electron microscopy to examine surface texture and
degradation and adjunct energy-dispersive analysis to detect the presence of
solder.

James Martin
Principal/Research Scientist
Orion Analytical, LLC
P.O. Box 550
Williamstown, MA 01267
www.orionanalytical.com

} i should clarify that the coin in question could not have been altered
} after 1900 so modern techniques are obviously out of the question. the
} issue becomes one of... if you wanted to alter a coin's date how would you
} have done it 100 years ago, and how could it be uncovered using modern
} analysis techniques.
}
} thanks!
} b-
} ----------------------------------------------
} Brian McIntyre
} Electron Microscopy Lab- River Campus
} Univ of Rochester
} Rochester, NY 14627
} 716-275-3058/4875
}





From: Caroline Schooley :      schooley-at-mcn.org
Date: Mon, 26 Jun 2000 13:03:39 -0700
Subject: Re: undergrad SEM use

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

You don't have to wait that long! Go to the Philadelphia meeting web page
search engine http://www.msa.microscopy.com/cgi-bin/M&M00Program.pl and
enter "teaching microscopy". Steve Barlow has organized an excellent
all-day session.

Caroline



Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Mon, 26 Jun 2000 21:16:07 +0100
Subject: Re: Electron Microscope Maintenance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Clearly, I have several vested interests here! As I am answering this
from home, I don't actually know as I write if you are specifically
concerned about a JEOL SEM, although I will check when I get to the
office tomorrow.

First, I would suggest you complain, with specifics, to the service
manager of the relevant company. If that is ineffective, direct your
complaint to the Salesperson covering your area - sales people,
perhaps, have a much stronger vested interest in happy customers. A
happy customer is possibly going to buy again in the future, an
unhappy customer makes their feelings known to others and can
influence future sales.

I hope you are not a JEOL user - if you are, please contact me
directly and I will attempt to resolve your problems.

If the SEM in question is an older instrument, independent service
companies will do a good job on all of the more routine problems. You
should also be able to resolve these yourself, in principle, with a
reasonable knowledge of vacuum technology and electronics.

I don't know where you might get training on SEM service - possibly Protrain?

The two main independent service companies that I am aware of
operating in the UK are ISS and EOS, both based in the Manchester
area but with engineers throughout most of the UK. Contact details
from sources on the web and Microscopy & Analysis. If you contect me
directly, I will forward you details.

Problems arise with newer instruments, especially PC-controlled ones.
Self-service and independent service will cover the basics but
without detailed training, circuit diagrams, software tools, it gets
very difficult to handle anything more than the basics.

Incidentally, this is one of the disadvantages for customers of
computer controlled EMs - it makes it easier for ALL manufacturers to
restrict service to approved service engineers. Not that this happens
just with scientific instruments - a similar issue exists with
cars/automobiles. I believe there is/has been a court case in the US
by independent auto-repair companies against the manufacturers to
force them to release service software.

Even on older instruments, major problems can be difficult to resolve
for anybody other than the manufacturer.

You will probably find that service via independents is cheaper than
via the manufacturers. Specialist parts will have to be purchased
from the manufacturer but even then, the overall cost will probably
be lower. However, I would add that I don't believe any EM
manufacturer is interested in providing anything other than
value-for-money service and, to a great extent, you do get what you
pay for - at least, you should! If you are paying less, overall, the
service you get is probably less - it may not be obvious, and you may
never see the difference but, in the long run, you will get what you
pay for.

You also need to keep in mind that the more comprehensive service
contracts provided by manufactuers contain a significant 'insurance'
element. Major components are expensive - with a comprehensive
service contract from the maufacturer, you will never know how
expensive but with service via an independent, you may find out!

Hope that is useful.

Best regards,
--
Larry Stoter
JEOL (UK) Ltd
Silver Court, Watchmead, Welwyn Garden City, AL7 1LT, United Kingdom
tel: +44-(0)1707-377117, fax: +44-(0)1707-373254, e-mail: larrys-at-jeoleuro.com




From: A. Greene :      ablue-at-io.com
Date: Monday, June 26, 2000 4:10 PM
Subject: question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

It would very much help to know what sort of an SEM you are using and at
what kV you are operating. I suspect you may have had a minor vacuum
accident which caused one of your "spray apertures" to be blown out of
position by inrushing air. Pull the liner tube (if your instrument has one)
and reposition all the apertures. It might not be a bad idea to do a good
cleaning job while inside the column.

Good luck.

Alex Greene
SCIENTIFIC INSTRUMENTATION SERVICES, INC.
PMB-499, 1807 West Slaughter Lane, Number 200
Austin, Texas 78748-6200
Phone: 512/282-5507 Fax: 512/280-0702

QUALITY ELECTRON MICROSCOPE REPAIR
-----Original Message-----
} From: Heejun Jeong {" hjjeong"-at-physics.purdue.edu} {Heejun Jeong {"
hjjeong"-at-physics.purdue.edu} }
To: Microscopy-at-sparc5.microscopy.com {Microscopy-at-sparc5.microscopy.com}





From: Harry Turner :      gen28t-at-happypuppy.com
Date: Mon, 26 Jun 2000 19:20:16 -0500
Subject: Between us... #3C9F

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


FREE E-COMMERCE WHEN YOU HOST WITH US!!

Tired of expensive e-commerce software, set up fees and leasing
contracts?
Here is the deal: Sign up for our E-Commerce Package and get a free
merchant account + a $200 cash coupon.

WE MEAN IT! THIS IS A REAL CASH COUPON! YOU PAY IN FACT $200 DOLLARS
LESS
DURING THIS SPECIAL PROMOTION. THIS OFFER IS ONLY GOOD FOR 7 DAYS.
DON'T
MISS THIS OPORTUNITY. NOBODY BEATS OUR PRICE!

While others charge you hundreds of dollars to get a merchant account
or
put you on a 48 months non-cancelable lease agreement we charge you
NOTHING for your merchant account when you sign up for one of our e
-commerce hosting plans. If you wish to stay with your current
hosting company or have already a merchant account our offer is even
better. We have 7 different E-Commerce plans to suit your individual
needs. We have a solution for U.S. beased merchants and international
merchants. Wherever you are on this planet, we get your online store
up and running within a few days.

Check it out first and make an informed decision. You have never seen
a
package deal like this before:

* Your own merchant account with one of the lowest rates in the
industry
* Real-Time software to accept VISA, MASTERCARD, AMEX, DISCOVER/NOVUS
,
DINERSCLUB/CARTE BLANCHE, JCB
* Direct deposit within 48 hrs into your checking account
* Shopping Cart store front software with an easy to use web based
interface
* Real-Time Credit Card Processing software
* Virtual terminal for phone/fax/mail orders
* Automated E-mail receipts to your clients
* Recurring billing feature with batch uploads
* Automatic batch closing
* Address verification system (AVS)
* Back office to 24/7 access account history
* 75 MB (megabytes) of disk space
* 30 GB (gigabyte) of data transfer per month
* 25 POP3 E-mail accounts
* Unlimited alias E-mail addresses
* Live web site statistics
* Unlimited FTP uploads
* Anonymous FTP
* CGI directory for your own scripts
* Site control panel
* Installation included
* Tech support included

All this and more when you sign up for our E-Commerce Hosting plan
for ONLY $69.95 per month and a one time set up fee of $199.00. That
's right. NO ADDITIONAL SET UP FEES or application fees for your
merchant account,
real-time software or shopping cart storefront. A one-stop E-Commerce
solution. And the best is:

NO LEASING, NO LONG TERM COMMITMENT. YOU CAN CANCEL ANYTIME.

In addition you get a FREE listing in our mall and FREE advertising
to
promote your store. We drive traffic to your site and help you to
become a
successful .com business.

REQUEST OUR FREE E-MAIL INFORMATION IMMEDIATELY! REMEMBER: THIS
SPECIAL
OFFER IS ONLY GOOD FOR 7 DAYS!

Please reply to:

mailto:tobk-at-mail.com?subject=INFO-PLEASE to receive our FREE e-mail
information package without obligations.

*********************************************************
Remove at mailto:gen28t-at-uymail.com?subject=remove
*********************************************************







From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 26 Jun 2000 20:51:29 -0700
Subject: Re: Power stability in microscopy and EELS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There are essentially two types of UPS systems. One acts in
a passive mode and switches to battery backup when power
fails. The other is an active mode which converts AC input
to DC and uses this to generate stable AC output.

Power conditioners are typically microprocessor-controlled units which
adjust taps on a transformer to maintain reasonably stable
output voltage. These can also include traps for spikes and
overvoltage conditions (Topaz for example).

The issue here is the stability of the basic potentials and currents
which are fed to or are delivered to the electron gun. In this
regard, I think that we are talking about very narrow margins of
input voltage variation and output voltages. I think that one will
have line sag no matter what--on occasion. Thus, the active
UPS units bypass these conditions and provide a constant,
stable AC supply to the SEM. Line conditioners are rated at
+/- 6% down to +/- 3%. Active UPS can do better.... but their
surge current capabilities are less than those of the conditioner units.

Consequently, there is not one, singularly perfect solution. But there are
options which the user should consider.

gg

At 12:29 AM 6/26/00, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Gordon Couger :      gcouger-at-couger.com
Date: Tue, 27 Jun 2000 04:34:29 -0500
Subject: Re: Power stability in microscopy and EELS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} From a practical point of view my experience with computer controlled UPS
systems is that cause more problems than they solve. They provide a
computer with good enough power in most cases. This power is not good
enough for equipment that is not designed for considerable voltage changes
and micro power losses.

Most of my experience comes from computer back up power supplies and a
project that was measuring sunlight. Since it was sunlight and we needed
fast response we did not have any low pass filter. When working on it in
the lab I used filtered DC since I was having problems with the 120 Hz
flicker of incandescent bulbs. I could see sags and glitches in the AC
power through the UPS systems even with the filters. It was good enough
for what I was doing but I could see a good deal difference in the
stability of the UPS and a battery.

I have worked with a few radio systems that use AC to charge a battery
bank and then power the radios from the batteries. These are far more
reliable than UPS systems.


Gordon
Gordon Couger gcouger-at-couger.com
Stillwater, OK www.couger.com/gcouger
405 624-2855 GMT -6:00

} From: "Dr. Gary Gaugler" {gary-at-gaugler.com}
}
} There are essentially two types of UPS systems. One acts in
} a passive mode and switches to battery backup when power
} fails. The other is an active mode which converts AC input
} to DC and uses this to generate stable AC output.
}
} Power conditioners are typically microprocessor-controlled units which
} adjust taps on a transformer to maintain reasonably stable
} output voltage. These can also include traps for spikes and
} overvoltage conditions (Topaz for example).
}
} The issue here is the stability of the basic potentials and currents
} which are fed to or are delivered to the electron gun. In this
} regard, I think that we are talking about very narrow margins of
} input voltage variation and output voltages. I think that one will
} have line sag no matter what--on occasion. Thus, the active
} UPS units bypass these conditions and provide a constant,
} stable AC supply to the SEM. Line conditioners are rated at
} +/- 6% down to +/- 3%. Active UPS can do better.... but their
} surge current capabilities are less than those of the conditioner units.
}
} Consequently, there is not one, singularly perfect solution. But there
are
} options which the user should consider.
}
} gg
}

} }
} } Hello all,
} }
} } Over time I saw some questions regarding the installation of a UPS or
some
} } power conditioner to improve the stability of all sorts of microscopy
} } equipment.
} } I understand that quite a few instruments are sensitive to the quality
of
} } the line power they get.
} } As I was working on EMC (electro magnetic compatibility) in a former
life,
} } this seems strange to me.
} } In Europe we have very strict EC-rules that guarantee that equipment
meets
} } its specifications even under severe conditions (including HF noise on
the
} } power, fluctuations in the power, high EM fields, temperature
variations
} } etc)
} }
} } I agree that microscopes are far more sensitive devices than, say,
} } videogames but does anyone know of the EMC rules for microscopes and
} } related equipment?
} } It seems strange that some people are using UPS's while especially
power
} } fluctuations can be easily overcome with a good design of the
} } electronics...
} }
} } Regards,
} }
} } Jo

}










From: donald j marshall :      dmrelion-at-world.std.com
Date: Tue, 27 Jun 2000 08:16:12 -0400 (EDT)
Subject: glass window thicknesses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am trying to locate a copy of a graph that shows the thicknesses of
(glass) windows that can support atmospheric pressure against a vacuum for
various radii of O ring support. Kaufman glass has one for high pressures
for several glass types but it doesn't cover the range of atmospheric
pressure and relatively small radii(1 to 3 cm). Pyrex, leaded glass, and
quartz are the three principal materials of interest. Appreciate any leads.

Don Marshall

Donald J. Marshall
Relion Industries
P.O. Box 12
Bedford, MA 01730
Ph: 781-275-4695
FAX: 781-271-0252
email dmrelion-at-world.std.com

Cathodoluminescence, mass spectroscopy, electron beam technology


"A weed is a flower out of place."

(Please note: Do not send email with attachments to this address. Instead, send it to donbarlen-at-aol.com. Thank you.)




From: Rahbari, Ramin :      RAMIN.RAHBARI-at-WL.com
Date: Tue, 27 Jun 2000 09:15:04 -0400
Subject: frogs' eggs and vesicles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You may want to contact
David Gard, Ph.D., Biology Department, University of Utah, Salt Lake City,
UT.
He has done a great deal of work with frog eggs.

good luck

Ramin Rahbari
Pfizer Global Research & Development
Worldwide Preclinical Safety
2800 Plymouth Road
Ann Arbor, MI 48105
Voice (734) 622-3383
Fax (734) 622-5001
Ramin.Rahbari-at-WL.COM


-----Original Message-----
} From: Mark West [mailto:mwest-at-ifcsun1.ifisiol.unam.mx]
Sent: Monday, June 26, 2000 3:37 PM
To: Microscopy-at-sparc5.microscopy.com


Hi,

A couple of weeks ago I wrote in witha problem trying to fix plant vesicle
preps for LR White inclusion and immunogold studies, without osmicating.
The pellets were disappearing during dehydration. I included an osmium
step (5, 30 or 60 minutes) and the pellets stopped disappearing, so I
guess there is so little protein that a formaldehyde/gluteraldehyde fix
doesn't manage to stabilize them. I'll have to check whether these
osmicated vesicles work in the immunogold reaction and check the effect
of a metaperiodate treatment to expose the proteins. Thanks for the advice
- another problem solved.

Someone brought us some frogs' eggs to fix. It looks like they have a hard
skin. The researcher is interested in extracting the nuclei (huge) which
also look like they're enclosed in a hard skin, and he wants to know
whether the extracted nuclei still retain some endoplasmic reticulum, so I
thought the best way would be to compare a whole egg with an extracted
nucleus. Any ideas/experience on how to fix this thing?

Thanks,

Mark






********************************************
Mark West,
Unidad de Microscopia Electronica,
(Electron Microscopy Unit)
Instituto de Fisiologia Celular,
Universidad Nacional Autonoma de Mexico,
04510 Mexico D.F.

tel (unidad/lab) *(525) 622 5610*
(casa/home) (525) 619 3020
Fax (525) 616 2282
********************************************





From: James M. Ehrman :      jehrman-at-mta.ca
Date: Tue, 27 Jun 2000 10:59:41 -0300
Subject: EDS on SEM w/o specimen airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,

We're in the process of acquiring an EDS system for our JEOL 5600 and
I'm curious about the experience of others with EDS systems on scopes
without specimen airlocks (column vents to exchange specimens). Any
thoughts/recommendations? I'm particularly concerned with contamination,

window integrity and the like. Are there any advantages/disadvantages of

Be vs. thin window detectors here?

Thanks,

Jim Ehrman

--

James M. Ehrman
Digital Microscopy Facility
Mount Allison University
Sackville, NB E4L 1G7
CANADA

phone: 506-364-2519
fax: 506-364-2505
email: jehrman-at-mta.ca
www: http://www.mta.ca/~jehrman






From: Dee Breger :      micro-at-ldeo.columbia.edu
Date: Tue, 27 Jun 2000 10:20:31 -0400 (EDT)
Subject: LM geological stain question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear colleagues,

I've received a query from a climatologist who wants to photograph fine
varves and/or laminations (~0.5 mm- 1 mm each layer) in carbonate (calcite)
rich silt (or marl) through a binocular light microscope. The samples are
flat, finely cut or polished sections that can be etched if necessary, but
he's really looking for a staining method that would enhance the difference
between the carbonate-rich layers and the clay-rich layers to make them
more usefully photogenic. Can anyone advise us on how to proceed to best
advantage?

Many thanks,
Dee




***************************************************************
Dee Breger
Mgr. SEM/EDX Facility
Lamont-Doherty Earth Observatory
61 Route 9W
Palisades, NY 10964 USA
T: 914/365-8640
F: 914/365-8155

http://www.ldeo.columbia.edu/micro
http://www.discovery.com/area/science/micro/micro1.html
http://www.lsc.org/antarctica/front.html
Journeys in Microspace (Columbia University Press, 1995)






From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Tue, 27 Jun 2000 10:13:20 -0500
Subject: Re: EDS on SEM w/o specimen airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have two different models of thin window detectors on two different
scopes. Our Hitachi 2460N does not have an airlock and is vented for every
sample change. Our JEOL 840A has an airlock, but does get the chamber
vented fairly often. We have only had one pin-hole develop in the detector
on our 840 in the 15-20 instrument-years that we have had the detectors.

Therefore, I don't think you would be gaining much, if anything, by
switching to a Be window. And you would be giving up that wonderful light
element capability. I would be hard pressed not to be able to detect C and
O. I would go for the thin window and be reasonably careful.

Warren S.

At 10:59 AM 6/27/2000 -0300, you wrote:

} Hi all,
}
} We're in the process of acquiring an EDS system for our JEOL 5600 and
} I'm curious about the experience of others with EDS systems on scopes
} without specimen airlocks (column vents to exchange specimens). Any
} thoughts/recommendations? I'm particularly concerned with contamination,
}
} window integrity and the like. Are there any advantages/disadvantages of
}
} Be vs. thin window detectors here?
}
} Thanks,
}
} Jim Ehrman
}
} --
}
} James M. Ehrman
} Digital Microscopy Facility
} Mount Allison University
} Sackville, NB E4L 1G7
} CANADA
}
} phone: 506-364-2519
} fax: 506-364-2505
} email: jehrman-at-mta.ca
} www: http://www.mta.ca/~jehrman
}





From: Vickie Frohlich :      frohlich-at-uthscsa.edu
Date: Tue, 27 Jun 2000 10:22:54 -0500
Subject: Re: Electron Microscope Maintenance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Could anyone recommend an independent U.K based company
} specialising in the repair and maintenance of electron microscopes ?
}
} Keen to hear comments ...


Information forwarded to list for Tributary Business Consulting:

I am an expense reduction professional. I have experience with equipment
maintenance contracts, and with Original Equipment Manufacture and
third-party equipment maintenance providers throughout the world. Please
contact me directly if you are interested.

Charles R. Frohlich CPA
President
Tributary Business Consulting
email: cfrohlich-at-satx.rr.com
phone: (210) 695-1364
fax: (210) 695-2141





From: George Lawton :      George.Lawton-at-email.swmed.edu
Date: Tue, 27 Jun 2000 10:03:18 -0500
Subject: SEM: Mounting Fly Heads

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


An investigator just brought me 6 samples - heads of flies. He was instructed to bring the entire body but he cuts the heads off when he does TEM. The investigator knows nothing about SEM. I have done alot of scanning of fly eyes but the body is attached.
My question is: how do I mount the heads on the stubs without the graphite or silver paint causing capillary action over the entire head? We will scan on a JEOL 840A after Au/Pd coating.

George Lawton
Chief Electron Microscopist
Molecular and Cellular Imaging Facility
UT Southwestern Medical Center at Dallas
5323 Harry Hines Blvd.
Dallas, Tx 75390-9039
Phone: 214-648-7291
Fax 214-648-6408
eMail: George.Lawton-at-email.swmed.edu




From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Tue, 27 Jun 2000 11:14:41 -0500
Subject: SEM: Mounting Fly Heads

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


George,

I would use one of the double-sided carbon sticky tabs available from
several vendors. They are good for small samples that would be buried in
liquids. If they don't provide enough contact area, you could very
carefully dab a conductive paint around the base of the mounted heads with a
pointy stick.

Cheers,
Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/




-----Original Message-----
} From: George Lawton [mailto:George.Lawton-at-email.swmed.edu]
Sent: Tuesday, June 27, 2000 10:03 AM
To: microscopy-at-sparc5.microscopy.com


An investigator just brought me 6 samples - heads of flies. He was
instructed to bring the entire body but he cuts the heads off when he does
TEM. The investigator knows nothing about SEM. I have done alot of
scanning of fly eyes but the body is attached.
My question is: how do I mount the heads on the stubs without the graphite
or silver paint causing capillary action over the entire head? We will scan
on a JEOL 840A after Au/Pd coating.

George Lawton
Chief Electron Microscopist
Molecular and Cellular Imaging Facility
UT Southwestern Medical Center at Dallas
5323 Harry Hines Blvd.
Dallas, Tx 75390-9039
Phone: 214-648-7291
Fax 214-648-6408
eMail: George.Lawton-at-email.swmed.edu




From: Doug Anderson :      danderson-at-schnabel-eng.com
Date: Tue, 27 Jun 2000 13:23:17 -0400
Subject: LM - Vibration Criteria -- Anybody know of them?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello --

I have dug through the archives of this fine list and found a bunch of
information on vibration and microscopy. I work for an engineering
consulting company, and part of our work is related to vibration of various
sorts (implosions, blasting, construction, pile driving, vehicular
vibration). A current project is for a medical facility that is expanding,
including operating rooms that have surgical microscopes. The facility is
very close to a freight line, and we are working with them to provide
mitigation measures before the facility is constructed.

My question to the list is, does anybody know of actual vibration criteria
in terms of amplitude (peak or RMS) and frequency for ANY microscopes. The
microscope manufacturers we have contacted indicate no research in the area,
which is surprising to me.

Our previous work with both human perception and machine sensitivity has
usually involved some specific vibration criterion, whether it be for
damage, annoyance, or whatever purpose. The discussion of the various
isolation tables on the list is interesting, including the innertubes and
tennis balls. They appear to be a "One isolator fits all" approach, which,
if it works, is also surprising to me. Any help or leads in this regard
would be greatly appreciated. Thanks

Doug Anderson

Douglas A. Anderson, PhD
Senior Consultant
Schnabel Engineering Associates (http://www.schnabel-eng.com)
510 East Gay Street
West Chester, PA 19380
Phone: 610 696-6066, Fax: 610 696-7771






From: Arrowood, Roy :      arrowood-at-utep.edu
Date: Tue, 27 Jun 2000 11:21:06 -0600
Subject: coin question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Brian, that's a very intriguing problem. What is the metal or alloy of the
coin? What is the particular date you are dealing with? Could a
counterfeiter have done something so simple as cutting out segments of a
pre-existing numeral to alter the date? If so, perhaps channelling would
reveal the coining deformation pattern below the missing segments.

Also, is it clear that the WHOLE COIN is not a counterfeit? Ordinarily, the
effort and cost to create a die set for counterfeiting a whole coin would
deter anyone from the effort, unless the coin is spectacularly valuable. On
the other hand, if someone had a genuine coin (or a carefully modified one)
to use as a pattern, it would not be too dreadfully difficult (even a
century ago) to make a ceramic mold directly, or a metal die indirectly,
from the pattern. Then you could produce "knock-offs" from that die. And
your "real McCoy" could be undamaged. This would take a clever, patient, and
very meticulous craftsman to pull it off, however.

====================================
Roy Arrowood, Associate Professor
Metallurgical and Materials Engineering
UTEP, El Paso, TX 79968-0520
(915)747-6934
NEW E-MAIL ADDRESS: arrowood-at-miners.utep.edu





From: Philip Oshel :      peoshel-at-facstaff.wisc.edu
Date: Tue, 27 Jun 2000 12:59:24 -0500
Subject: Re: SEM: Mounting Fly Heads

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This worked for amphipod maxilla, which are minute or smaller and
setose, so it should work for fly heads:
1) mount on double-sticky carbon-conductive tabs; cut these from a
sheet, or buy as stub-sized tabs
2) sharpen toothpicks or other sticks, some to a point and others to
a pen nib (with or without the slot found in a fountain pen nib)
3) place a small blob of Ag paint at the edge of the stub connecting
the surface of the sticky tab to the metal of the stub; I prefer Ag
paint dissoved in methylethylketone
4) use the stick to draw the paint from the blob to the fly head
(etc.); connect this to a part of the head that you don't care about;
draw a circle around the head just far enough not to touch the head
-- this creates a shorter conductive path from the specimen to the Ag
paint

Careful use of the sticks and paint will allow you to use Ag paint on
very small, hairy specimens. The paint can be allowed to partially
dry (solvent to evaporate) to thicken it, but this may just form a
skin on the surface, and does lead to stringing of the partly dry
paint which can be very annoying.

Phil

} An investigator just brought me 6 samples - heads of flies. He was
} instructed to bring the entire body but he cuts the heads off when
} he does TEM. The investigator knows nothing about SEM. I have done
} alot of scanning of fly eyes but the body is attached.
} My question is: how do I mount the heads on the stubs without the
} graphite or silver paint causing capillary action over the entire
} head? We will scan on a JEOL 840A after Au/Pd coating.
}
} George Lawton
} Chief Electron Microscopist
} Molecular and Cellular Imaging Facility
} UT Southwestern Medical Center at Dallas
} 5323 Harry Hines Blvd.
} Dallas, Tx 75390-9039
} Phone: 214-648-7291
} Fax 214-648-6408
} eMail: George.Lawton-at-email.swmed.edu

--
}}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
Philip Oshel
Supervisor, AMFSC and BBPIC
Dept. of Animal Health and Biomedical Sciences
University of Wisconsin
1656 Linden Drive
Madison, WI 53706-1581
voice: (608) 263-4162
fax: (608) 262-7420 (dept. fax)




From: Richard E. Edelmann :      edelmare-at-casmail.muohio.edu
Date: Tue, 27 Jun 2000 15:33:59 -0500
Subject: Microscopy and Molecular Biology Postdoctoral Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




{color} {param} 0100,0100,0100 {/param} Please respond directly to Dr. David Pennock at {/color} pennocdg-at-muohio.edu. {bold} {FontFamily} {param} TIMES {/param} {/bold} {color} {param} 0100,0100,0100 {/param} {FontFamily} {param} Arial {/param}



=======================================================


{bold} {bigger} {bigger} Microscopy and Molecular Biology Postdoctoral
Position

Department of Zoology

Miami University

Oxford, OH {/bold} {FontFamily} {param} TIMES {/param}


{FontFamily} {param} Arial {/param} A two-year postdoctoral position is available for
investigating the role of inner arm dyneins in ciliary motility.
We have seven different inner arm dynein heavy chain
genes cloned from {underline} Tetrahymena {/underline} {underline} thermophila {/underline} and are in
the process of creating knockout mutations in those
genes. The successful applicant will be involved in all
aspects of the work, including creation of the knockout
mutants and analysis of the phenotypes. The successful
candidate should be experienced in light and electron
microscopy and willing to learn molecular biology and
some protein biochemistry (preferred) or experienced in
molecular biology and protein biochemistry and willing to
learn light and electron microscopy. Send letter,
curriculum vitae, and reference letters to:


David Pennock

Department of Zoology

Miami University, Oxford, OH 45056


Email: pennocdg-at-muohio.edu {FontFamily} {param} TIMES {/param} {smaller} {smaller}



{/x-rich}



From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Tue, 27 Jun 2000 15:38:04 -0400
Subject: SEM: Mounting Fly Heads

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


HI

Drill small holes in stubs and glue a dress making pin in each hole spike
upwards. Coat this in a sputter coater.

Push each insect head onto a pin and sputter coat at a specimen-target
distance of 5cms.

This technique is used by a number of our clients.

Steve Chapman
Senior Consultant Protrain
For consultancy and professional training in EM world wide.
www.emcourses.com
Tel 44+ 1280 814774 Fax 44+ 1280 814007




From: Matthew Lynn :      mlynn-at-miami.edu
Date: Tue, 27 Jun 2000 16:54:06 -0400
Subject: RE: LM geological stain question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Dee,

I've done a little work with varves...let me start by saying that typically
this imaging is done using X-rays or by making conventional thin sections for
petrographic microscopy. That said, I vaguely recall staining carbonate thin
sections with Alizarin Red S. There is a different stain for dolomite. I'm
not sure how well the technique can be applied to large sections....are they
still wet? Dehydrated? Vacuum impregnated? Let me know if this is on the
right track and I will hunt for some references.

Matt

Matthew J. Lynn, Ph.D.
Center for Advanced Microscopy
University of Miami
(305)284-4736
mlynn-at-miami.edu


On Tuesday, June 27, 2000 10:21 AM, Dee Breger [SMTP:micro-at-ldeo.columbia.edu]
wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear colleagues,
}
} I've received a query from a climatologist who wants to photograph fine
} varves and/or laminations (~0.5 mm- 1 mm each layer) in carbonate (calcite)
} rich silt (or marl) through a binocular light microscope. The samples are
} flat, finely cut or polished sections that can be etched if necessary, but
} he's really looking for a staining method that would enhance the difference
} between the carbonate-rich layers and the clay-rich layers to make them
} more usefully photogenic. Can anyone advise us on how to proceed to best
} advantage?
}
} Many thanks,
} Dee
}
}
}
}
} ***************************************************************
} Dee Breger
} Mgr. SEM/EDX Facility
} Lamont-Doherty Earth Observatory
} 61 Route 9W
} Palisades, NY 10964 USA
} T: 914/365-8640
} F: 914/365-8155
}
} http://www.ldeo.columbia.edu/micro
} http://www.discovery.com/area/science/micro/micro1.html
} http://www.lsc.org/antarctica/front.html
} Journeys in Microspace (Columbia University Press, 1995)
}





From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Tue, 27 Jun 2000 15:01:49 -0700
Subject: Re: EDS on SEM w/o specimen airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Jim,
We have a Kevex Quantum light element EDX on a Hitachi that vents the whole
chamber and it has been no problem, but I am told that the rate of venting
can make quite a difference. I bought our EDX detector used and the previous
owner claimed to have replaced the window every year for the five years that
he had it. I have never had to replace the window and that is in a student
lab. This particulars SEM takes over a minute to vent and makes no
discernable sound, whereas my other SEM vents in ten seconds with an audible
hiss. It, fortunately, has a Be-window detector, which is definitely
tougher. If your SEM vents quickly, you might try to put some sort of
restricter or filter on the vent to clean and slow it down.
At 10:59 AM 6/27/00 -0300, you wrote:
}
} Hi all,
}
} We're in the process of acquiring an EDS system for our JEOL 5600 and
} I'm curious about the experience of others with EDS systems on scopes
} without specimen airlocks (column vents to exchange specimens). Any
} thoughts/recommendations? I'm particularly concerned with contamination,
}
} window integrity and the like. Are there any advantages/disadvantages of
}
} Be vs. thin window detectors here?
}
} Thanks,
}
} Jim Ehrman
Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca





From: Jane Fagerland :      jane.a.fagerland-at-abbott.com
Date: Tue, 27 Jun 2000 17:22:05 -0500
Subject: Research Cellular/Molecular Biologist - job announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A colleague of mine has asked me to post the follwoing job announcement for
an opening in our Cancer Research area. Please send resumes to the address
at the bottom of the ad, or visit the website, which is also at the bottom of
the ad. Please do not contact me directly concerning this position. Thanks!

Jane A. Fagerland, Ph.D.
Dept. Microscopy and Microanalysis
Abbott Laboratories

******************************************************************************
******

ABBOTT LABORATORIES
PHARMACEUTICAL PRODUCTS DIVISION
Abbott Park, Illinois

Abbott Laboratories is a global diversified company dedicated to the
discovery, development, manufacture and marketing of health care products and
services. Around the world, the company,s 54,000+ employees are committed to
improving people,s lives by providing cost-effective health care
technologies. We are interested in interviewing candidates for the following
opportunity:

Research Cellular/Molecular Biologist

Participate in understanding the mechanism of action of novel anti-cancer
drugs with our in vivo oncology team. Areas of research include tumor
growth, angiogenesis, apoptosis and metastasis. Major responsibilities will
include immunohistochemistry, immunocytochemistry, in situ hybridization and
cellular analysis of tumor sections. Requires a Master,s degree with 3-5
years industrial or academic research experience. Must have experience in
immunohistochemistry, immunocytochemistry, in situ hybridization, tissue
preparations, general histology and cell cycle analysis. Experience with
cellular in vitro assays, tissue culture, western analysis and some molecular
biology techniques are desired. Requires good interpersonal and
communication skills.

Please send resumes to: 100 Abbott Park Road, D-583, AP9A, Abbott Park,
Illinois 60064 Attn. Job Code 2K-KDA3422. Or visit our web site at
www.abbott.com

An EOE, we are committed to employee diversity




From: James Talbot :      jptkt-at-flash.net
Date: Tue, 27 Jun 2000 18:08:58 -0500
Subject: LM geological stain question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dee-

Before staining, try using a UV light source to see if the calcite
fluoresces - can fluoresce red to pink, orange, white, yellow or blue.

If you need to stain, use Alizarin Red S for staining. If it turns a deep
red color, it's most likely calcite. Procedures for this can be found in
"Laboratory Handbook of Petrographic Techniques" by Charles S. Hutchinson,
1974, John Wiley and Sons. Or you can have a thin section preparation lab
do the staining for you - this will be much easier since they are set up to
do this routinely.

Cheers,
James Talbot

K/T GeoServices, Inc.
X-ray diffraction petrologic studies
visit my web site at http://www.ktgeo.com
Argyle, TX, USA, (214) 403-6342

} ----- Original Message -----
} From: Dee Breger {micro-at-ldeo.columbia.edu}
} To: {microscopy-at-sparc5.microscopy.com}
} Sent: Tuesday, June 27, 2000 9:20 AM
} Subject: LM geological stain question
}
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Dear colleagues,
} }
} } I've received a query from a climatologist who wants to photograph fine
} } varves and/or laminations (~0.5 mm- 1 mm each layer) in carbonate
} (calcite)
} } rich silt (or marl) through a binocular light microscope. The samples
are
} } flat, finely cut or polished sections that can be etched if necessary,
but
} } he's really looking for a staining method that would enhance the
} difference
} } between the carbonate-rich layers and the clay-rich layers to make them
} } more usefully photogenic. Can anyone advise us on how to proceed to
best
} } advantage?
} }
} } Many thanks,
} } Dee
} }
} }
} }
} }
} } ***************************************************************
} } Dee Breger
} } Mgr. SEM/EDX Facility
} } Lamont-Doherty Earth Observatory
} } 61 Route 9W
} } Palisades, NY 10964 USA
} } T: 914/365-8640
} } F: 914/365-8155
} }
} } http://www.ldeo.columbia.edu/micro
} } http://www.discovery.com/area/science/micro/micro1.html
} } http://www.lsc.org/antarctica/front.html
} } Journeys in Microspace (Columbia University Press, 1995)
} }
} }
} }
} }
}





From: Gib Ahlstrand :      giba-at-puccini.cdl.umn.edu
Date: Tue, 27 Jun 2000 18:55:06 -0500 (CDT)
Subject: Re: SEM: Mounting Fly Heads

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


George,

I've mounted small insects similar to what you now have using the following
method:

1. Get some dissecting pins, preferably, or any very thin, rigid pin or wire and
cut about 1/4 inch long.

2. Mount it vertically on an SEM stub using carbon paint, or any quick drying
glue, and let it dry completely, put in warming oven to speed it up, or use hair
dryer.

3. Position the fly heads upsidedown on clean surface, like lens tissue or glass
petri dish.

4. Put a fresh drop of carbon paint, or glue, on a nearby surface, pick up the
SEM stub with stub handling forceps, dip the tip of the pin into the fresh paint
or glue just enough to get a tiny little blob on the pin head, then under a
dissecting or low power scope touch the tip of the pin to the back surface of
the fly head. It should stick and you can pick it up at that point, place right
side up and let dry completely.

5. Coat in vacuum evaporator or sputter coater as usual. After coating run, vent
gas into chamber slowly to not disturb delicate sample.

Its a bit of a prep, but you also get the fly head off the stub surface, little
or no background junk in the view. Good luck!

Gib



Responding to the message of {s9587bef.020-at-mednet.swmed.edu}
from "George Lawton" {George.Lawton-at-email.swmed.edu} :
}
} An investigator just brought me 6 samples - heads of flies. He was
} instructed to bring the entire body but he cuts the heads off when he does
} TEM. The investigator knows nothing about SEM. I have done alot of scanning
} of fly eyes but the body is attached.
} My question is: how do I mount the heads on the stubs without the graphite
} or silver paint causing capillary action over the entire head? We will scan
} on a JEOL 840A after Au/Pd coating.
}
} George Lawton
} Chief Electron Microscopist
} Molecular and Cellular Imaging Facility
} UT Southwestern Medical Center at Dallas
} 5323 Harry Hines Blvd.
} Dallas, Tx 75390-9039
} Phone: 214-648-7291
} Fax 214-648-6408
} eMail: George.Lawton-at-email.swmed.edu
}
}
} .


Gib Ahlstrand
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN. USA. 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.cdl.umn.edu
http://biosci.umn.edu/MIC/consortium.html





From: Ron Doole :      ron.doole-at-materials.oxford.ac.uk
Date: Wed, 28 Jun 2000 08:11:38 +0100 (BST)
Subject: Re: EDS on SEM w/o specimen airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,

Just a reminder while we are on the subject of thin window detectors -
as well as controlling the gas flow on venting also make sure that the
chamber can not overpressure.
On SEMs the chamber door should be able to open freely as soon as it
reaches atmospheric pressure. On TEMs an aperture mechanism or similar
should be freed so that it can release at atmospheric pressure.

Ron



On Tue, 27 Jun 2000, Mary Mager wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear Jim,
} We have a Kevex Quantum light element EDX on a Hitachi that vents the whole
} chamber and it has been no problem, but I am told that the rate of venting
} can make quite a difference. I bought our EDX detector used and the previous
} owner claimed to have replaced the window every year for the five years that
} he had it. I have never had to replace the window and that is in a student
} lab. This particulars SEM takes over a minute to vent and makes no
} discernable sound, whereas my other SEM vents in ten seconds with an audible
} hiss. It, fortunately, has a Be-window detector, which is definitely
} tougher. If your SEM vents quickly, you might try to put some sort of
} restricter or filter on the vent to clean and slow it down.
} At 10:59 AM 6/27/00 -0300, you wrote:
} }
} } Hi all,
} }
} } We're in the process of acquiring an EDS system for our JEOL 5600 and
} } I'm curious about the experience of others with EDS systems on scopes
} } without specimen airlocks (column vents to exchange specimens). Any
} } thoughts/recommendations? I'm particularly concerned with contamination,
} }
} } window integrity and the like. Are there any advantages/disadvantages of
} }
} } Be vs. thin window detectors here?
} }
} } Thanks,
} }
} } Jim Ehrman
} Regards,
} Mary
}
} Mary Mager
} Electron Microscopist
} Metals and Materials Engineering
} University of British Columbia
} 6350 Stores Road
} Vancouver, B.C. V6T 1Z4
} CANADA
} tel: 604-822-5648
} e-mail: mager-at-interchg.ubc.ca
}
}
}

===========================================================================
Mr. Ron Doole e-mail ron.doole-at-materials.ox.ac.uk
Department of Materials, phone +44 (0) 1865 273701
University of Oxford, fax +44 (0) 1865 283333
Parks Road.
Oxford. OX1 3PH. UK.
============================================================================





From: Stephan Thiberge :      thiberge-at-wicc.weizmann.ac.il
Date: Wed, 28 Jun 2000 15:50:11 +0300
Subject: cathodoluminescent detector

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I want to buy a cathodoluminescent detector for a Scanning Electron Microscope JEOL JSM 6400.

I am looking for a company which has this product.

If you know such a company I would greatly appreciate your help.

Thanks
Stephan







From: COURYHOUSE-at-aol.com
Date: Wed, 28 Jun 2000 11:25:16 EDT
Subject: Re: Change of e-mail address and reminder of museum want list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


OK thanks for the change of address, as usual we are looking for:
books on microscopes
catalogs of microscopes
microscope instruction manuals
books on application of microscopes either in biological or material sciences
.. anything! Optical, em sem, afm, etc etc etc.

-----
items and literature relating to any form of electrical communication and
engineering of thereof (yes telephone, telegraph, early wireless telegraphy
----
artifacts and literature on radio and TV broadcasting
---
Radar and radar countermeasures
-----
cryptanalysis
--------

thanks in advance as always
Ed Sharpe archivist for SMECC




From: Tom Phillips :      PhillipsT-at-missouri.edu
Date: Wed, 28 Jun 2000 12:20:51 -0500
Subject: scanning EM negative question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have used my Umax Powerlook III flatbed scanner with transparency
adapter many times to get high quality scans of my TEM negatives but
today it is driving me crazy. I have some great gold labeling of
pretty electron dense granules. On the negative I can clearly see
the gold against the granule matrix but when I scan, the 256 grey
levels compresses all that info into one level of black and therefore
I can't discern the gold. any scan gurus out there who can help? I
have tried playing with the gamma without much luck. thanks in
advance, tom

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility

3 Tucker Hall
Division of Biological Sciences
University of Missouri
Columbia, MO 65211-7400
(573)-882-4712 (voice)
(573)-882-0123 (fax)




From: Greg Erdos :      gwe-at-biotech.ufl.edu
Date: Wed, 28 Jun 2000 14:20:13 -0400
Subject: Re: scanning EM negative question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tom,
Have you tried telling the scanner that the image is a positive
transparency rather than a negative? And then reverse the contrast in
photoshop rather than with the scanner software.
Greg

At 12:20 PM 06/28/2000 -0500, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Gregory W. Erdos, Ph.D.
Assistant Director, Biotechnology Program
PO Box 110580
University of Florida
Gainesville, FL 32611

Ph. 352-392-1295 Fax: 352-846-0251




From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Wed, 28 Jun 2000 10:32:35 -1000 (HST)
Subject: Re: scanning EM negative question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


we have a Epson 1600, sometimes this problem happens too. my solution is to
scan the negative in positive mode first then do a inversion after that.
Hope this helps.

Hao Li

----- Original Message -----
} From: "Tom Phillips" {PhillipsT-at-missouri.edu}
To: {Microscopy-at-sparc5.microscopy.com}
Sent: Wednesday, June 28, 2000 1:20 PM


Hi, Thomas

I find it best to scan a negative in as a positive transparency at 6
million colors (rather than the 256 greys you are using), and then reverse
it. This gives me the best tonal range.

Aloha,
Tina


} } I have used my Umax Powerlook III flatbed scanner with transparency
} } adapter many times to get high quality scans of my TEM negatives but
} } today it is driving me crazy. I have some great gold labeling of
} } pretty electron dense granules. On the negative I can clearly see
} } the gold against the granule matrix but when I scan, the 256 grey
} } levels compresses all that info into one level of black and therefore
} } I can't discern the gold. any scan gurus out there who can help? I
} } have tried playing with the gamma without much luck. thanks in
} } advance, tom
} }

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************





From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Wed, 28 Jun 2000 21:08:13 +0100
Subject: Re: cathodoluminescent detector

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Oxford Instruments supply a wide range of CL detection systems, from
low cost broad-band detectors to multiple-band, UV, IR and
spectroscopic systems. As far as I am aware, they are the only
company supplying such a wide range of CL systems - if there are
others, sorry, and I would also like to know!

Regards
--
Larry Stoter
JEOL (UK) Ltd
Silver Court, Watchmead, Welwyn Garden City, AL7 1LT, United Kingdom
tel: +44-(0)1707-377117, fax: +44-(0)1707-373254, e-mail: larrys-at-jeoleuro.com




From: Tom Phillips :      PhillipsT-at-missouri.edu
Date: Wed, 28 Jun 2000 15:42:05 -0500
Subject: scanning EM negative problem solved

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks for all the suggestions. We did indeed find that scanning in
the color mode (which is 14 bit) solved the problem. I don't know
why UMax has their greyscale image only go to 256 colors. I will
ultimately incorporate this in a lecture when i am teaching image
analysis since it clearly shows the limitations of 8 bit images.
once again, thanks.

tom

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility

3 Tucker Hall
Division of Biological Sciences
University of Missouri
Columbia, MO 65211-7400
(573)-882-4712 (voice)
(573)-882-0123 (fax)




From: Jonathan Barnard :      Jonathan.Barnard-at-angstrom.uu.se
Date: Wed, 28 Jun 2000 23:53:07 +0200
Subject: Re: SEM & CL

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Stephan,
I know that Oxford instruments can supply you with the necessary equipment
to do CL in the SEM. I am not sure whether this is standard equipment or
whether it is special order, but I have seen it done. The equipment usually
involves a parabolic mirror and a light-pipe with the relevant detector on
the end of this. Then there is the software .etc. There is a contact email
address on their web page:

http://www.oxford-instruments.com/

give them a try and see what they say.

Regards,
Jonathan



********************************************************
Dr Jonathan Barnard

Analytical Materials Physics
The Angstrom Laboratory, Uppsala University
P O Box 534, SE-751 21 Uppsala, Sweden
Phone: +46-(0)18-4716838 Fax: +46-(0)18-500131
Phone: Microscope room +46 18 471 6365
http://www.angstrom.uu.se/analytical/home.html
********************************************************





From: jmkrupp-at-cats.ucsc.edu (Jon Krupp)
Date: Wed, 28 Jun 2000 15:10:15 -0700
Subject: Have you tried MDS 100?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi:

Has anyone tried Kodak's MDS 100 imaging system? I have a user who needs a
quick documentation system and this might work for her. She is using NIH
Image on a compound scope in our lab and she wants to become independent
and do the work in her own lab, and maybe at other locations.

As an alternative, anyone with a Nubus frame grabber, ie Scion LG3 or
equiv., that would sell it to her to use with NIH Image?

Jonathan Krupp
Microscopy & Imaging Lab
University of California
Santa Cruz, CA 95064
(831) 459-2477
jmkrupp-at-cats.ucsc.edu






From: Caroline Schooley :      schooley-at-mcn.org
Date: Wed, 28 Jun 2000 16:52:54 -0700
Subject: Re: scanning EM negative question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} I have used my Umax Powerlook III flatbed scanner with transparency
} adapter many times to get high quality scans of my TEM negatives but
} today it is driving me crazy. I have some great gold labeling of
} pretty electron dense granules. On the negative I can clearly see
} the gold against the granule matrix but when I scan, the 256 grey
} levels compresses all that info into one level of black and therefore
} I can't discern the gold. any scan gurus out there who can help? I
} have tried playing with the gamma without much luck. thanks in
} advance, tom
}
} Thomas E. Phillips, Ph.D.
} Associate Professor of Biological Sciences
} Director, Molecular Cytology Core Facility
}
Tom -

You can get a really good book on scanner use at www.scantips.com or
download it as a pdf,if you have the patience for 218 pp.

Caroline

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html






From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Thu, 29 Jun 2000 16:05:07 GMT+1200
Subject: JEOL 840A cooling water flow

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi

Can anyone advise what the water flows should be in the three
parallel branches (OL; DPs; and power boards) of the 840A)
The manual spec is } 5 l/min total, but I need to know particularly
what the flow should be thru the OL.

TIA

rtch

ps replies welcome from JEOL


Ritchie Sims Phone : 64 9 3737599 ext 7713
Department of Geology Fax : 64 9 3737435
The University of Auckland email : r.sims-at-auckland.ac.nz
Private Bag 92019
Auckland
New Zealand




From: Stafford McKnight :      s.mcknight-at-ballarat.edu.au
Date: Thu, 29 Jun 2000 15:49:10 +1000
Subject: Cathodoluminescence detector

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Stephan,

We have an Oxford MiniCL detector attached to a JEOL 6300 -
same column as the 6400. This is an inexpensive detector which
suits our needs but there are other suppliers.
No matter who you obtain one from, give some thought to what
ports are free and to your analytical working distance (presume you
have EDS or WDS). On the 6400 this is most likely 15mm - some
ingenuity from Oxford or other supplier will be needed to clear the
pole-piece eg an inclined configuration. It would be worth it - rather
than have to alter the sample height for quant analysis.

Regards

Stafford
Stafford McKnight
Geology & Metallurgy
University of Ballarat
ph 03 53279262
fax 03 53279144




From: DavidSu-at-aol.com
Date: Thu, 29 Jun 2000 02:06:30 EDT
Subject: SEM Positions Open

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Accurel Systems International is looking for SEM Analysts: junior to
experienced technicians and engineers

Responsibilities include all aspects of commercial SEM and EDX analysis
(instrument operation, sample preparation, customer contact and follow-up,
data interpretation etc.).

Knowledge of IC device structure, IC fabrication and packaging a plus.

2+ years hands on experience in the semiconductor industry preferred.

To find out more please check our web site: www.accurel.com

SEM candidates should send resume with references to Regina Campbell:
reginac-at-accurel.com or FAX 408-737-3916.




From: richard.beanland-at-gecm.com
Date: Thu, 29 Jun 2000 09:15:09 +0000 (GMT)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe Microscopy richard.beanland-at-gecm.com





From: richard.beanland-at-gecm.com
Date: Thu, 29 Jun 2000 11:04:31 +0000 (GMT)
Subject: Position available at Caswell Technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopist/engineer:
Structural analysis department,
Marconi Caswell Ltd.,
Towcester,
Northants NN12 8EQ
UK

Job description:
You will form part of a team of specialists in analytical techniques, providing support to the rapidly growing III-V optoelectronic and microwave device fabrication plant at Caswell Technology. We perform a wide range of activities, including optical microscopy, SEM, TEM, EDX, X-ray techniques and scanning probe microscopies. You will be expected to be able to be competent in several of these techniques and have/develop a degree of expertise in one or more of them. Experience of semiconductor device fabrication techniques would be an advantage.

The post is available immediately.

More details of our company are available at
http://www.caswelltechnology.com/

Please contact me if you are interested or would like any further information.

==============================================================
Richard Beanland
Caswell Technology,
Caswell,
Towcester,
Northants NN12 8EQ

e-mail richard.beanland-at-gecm.com
Tel. +44 1327 356363
Fax. +44 1327 356398
http://www.caswelltechnology.com/
==============================================================





From: Nestor J. Zaluzec :      zaluzec-at-aaem.amc.anl.gov
Date: Wed, 28 Jun 2000 11:23:51 -0500
Subject: Administrivia: Change of Address Notices

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Folks

Don't broadcast change of address notices to the Listserver.
If you want to change your subscription address, just send
a unsubscribe message, followed by a subscribe message...

BUT SEND IT TO THE ADMINISTRATIVE ADDRESS!!!!

LISTSERVER-at-MSA.MICROSCOPY.COM

don't send them to the posting address

(Microscopy-at-MSA.MICROSCOPY.COM)

Nestor
Your Friendly Neighborhood SysOp

==================================================================
Nestor J. Zaluzec
Materials Science Division
Building 212
Argonne National Lab
9700 S. Cass Ave
Argonne, Illinois 60439 USA
Tel: 630-252-7901, Fax: 630-252-4798
Email: Zaluzec-at-aaem.amc.anl.gov
==================================================================
TPMLab: http://tpm.amc.anl.gov
MMSite: http://www.amc.anl.gov
==================================================================

The box said "This program requires Win 95/98/NT or better..." so I bought
a G3 Mac

==================================================================






From: richard black :      m02jmy00-at-cwcom.net
Date: Thu, 29 Jun 2000 07:58:15 -0500
Subject: Printer information please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I run a microscopy unit which is completely digital. I have a problem
with printing out images. I am planning to upgrade cameras to 10 or 12
bit megapixel units, but am already restricted by printers. Laser
printers seem to deal with greyscales by halftoning at 100-150dpi which
is unsatisfactory for small prints. Are there any true greyscale
printers available, that manage better than 300dpi and better than 256
grey levels? Is there any advantage in having more than 256 grey
levels? Sorry if this is a bit dim, but there are so many self styled
experts who give conflicting opinions I end up totally confused.
Thank you.
Richard Black
Nottingham, England
richardblack-at-cwcom.net






From: Tim Lyden, Ph.D. :      lyden.11-at-osu.edu
Date: Thu, 29 Jun 2000 09:12:54 -0400
Subject: Scan site

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is my first message to the list... Caroline, I looked at Scan site and
would like to try PDF format... but I can seem to locate the pathway to
download? Any thoughts or pointers?

Thanks for the suggestion it looks like a great site.
Tim Lyden, Ph.D.
Research Scientist

Departments of Internal Medicine/Immunology and Physiology/Cell Biology
Ohio State University
Davis Medical Research Center
480 W 9th Ave.
Columbus, Ohio 43210


Phone: 614-293-4867
Fax: 614-293-5631





From: Alesia Darling :      adarling-at-coe.eng.ua.edu
Date: Thu, 29 Jun 2000 09:10:34 CST
Subject: Tunneling Electron Microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to purchase a "Tunneling Electron Microscope".

I have been unable to locate a supplier. If you know of one please
let me know. Thanks!
Alesia White Darling
Secretary, Dr. Amnon Katz
University of Alabama
Aerospace Engineering and Mechanics
Box 870280/205 Hardaway Hall
Tuscaloosa, Alabama 35487-0280
ph: 205-348-8525
fax: 205-348-7240 or 205-348-2094




From: Soumitra Ghoshroy :      ghoshroy-at-nmsu.edu
Date: Thu, 29 Jun 2000 07:13:50 -0700
Subject: Re: scanning EM negative question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I have found that the gold does not show up well when you scan the negative
itself. So for gold to show up in the scanned image, I usually make one
print and scan the print. This works quite well. I guess we will still have
to work in darkrooms for quite sometime even though the digital imaging is
becoming popular.

Good luck,

Soumitra

}
}
} I have used my Umax Powerlook III flatbed scanner with transparency
} adapter many times to get high quality scans of my TEM negatives but
} today it is driving me crazy. I have some great gold labeling of
} pretty electron dense granules. On the negative I can clearly see
} the gold against the granule matrix but when I scan, the 256 grey
} levels compresses all that info into one level of black and therefore
} I can't discern the gold. any scan gurus out there who can help? I
} have tried playing with the gamma without much luck. thanks in
} advance, tom
}
} Thomas E. Phillips, Ph.D.
} Associate Professor of Biological Sciences
} Director, Molecular Cytology Core Facility
}
} 3 Tucker Hall
} Division of Biological Sciences
} University of Missouri
} Columbia, MO 65211-7400
} (573)-882-4712 (voice)
} (573)-882-0123 (fax)




*****************************************************************
Soumitra Ghoshroy Ph.D.
Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3600
Fax: 505-646-5665
e-mail: ghoshroy-at-nmsu.edu
http://confocal.nmsu.edu/eml






From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Thu, 29 Jun 2000 11:21:33 -0500
Subject: Re: Printer information please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


As I recall, the human eye is hard pressed to distinguish more than 6-bit
grayscale (64 levels) unless the shades are right next to each other, which
they might be in an image. Therefore, printing at 256 gray levels should be
more than adequate.

The 10 or 12-bit cameras will still be handy in storing the extra
information in your images. You will be able to do some gray scale
manipulation to highlight the darker or lighter regions of your image. But
when it comes to print time, you will be throwing out all but about 8 bits.

You rightly say that halftoning is a problem on small prints. You need a
16x16 dot cell to render 256 gray levels in halftone. For 1200 dpi
printing, that means that you can achieve
75 pixels per inch (not very high). That means a 1024 pixel image would
require 13.65 inches to show all that resolution. If the image is printed
out smaller, then either grayscale or pixel resolution would be sacrificed.
(Printing at 64 gray levels would require a 8x8 cell and could be done at
150 pixels per inch so that the image could fit into 6.8 inches.)

I don't have a dye sub printer here, but their specs would probably meet
your needs. You would want to see how well they can render the grayscale.
Since they do not halftone, the dpi spec would be the same as the pixel per
inch spec. A 300 dpi printer should be able to render full resolution on a
1024 pixel image in 3.4 inches. But check out the grayscale performance to
see if it is adequate.

Warren S.

At 07:58 AM 6/29/2000 -0500, you wrote:

} I run a microscopy unit which is completely digital. I have a problem
} with printing out images. I am planning to upgrade cameras to 10 or 12
} bit megapixel units, but am already restricted by printers. Laser
} printers seem to deal with greyscales by halftoning at 100-150dpi which
} is unsatisfactory for small prints. Are there any true greyscale
} printers available, that manage better than 300dpi and better than 256
} grey levels? Is there any advantage in having more than 256 grey
} levels? Sorry if this is a bit dim, but there are so many self styled
} experts who give conflicting opinions I end up totally confused.
} Thank you.
} Richard Black
} Nottingham, England
} richardblack-at-cwcom.net





From: Lehman, Ann :      Ann.Lehman-at-trincoll.edu
Date: Thu, 29 Jun 2000 14:02:38 -0400
Subject: Balzers Contact Info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think this query was posted recently but I haven't been able to dredge up
the replies in the list archives - too recent, maybe.

I'm looking for contact info for a colleague, for Balzers or Bal-tec. He
inherited a B. carbon evaporator model CED 030 (also: 3U-G03-751/132). Can
anyone help out with either contact info for the company, or a copy of the
manual?? We will pay copy/ship charges.

Thanks, much.

Ann Hein Lehman
Manager, EM Facility
Trinity College
Hartford CT 06106
v. 860-297-4289
f. 860-297-2538
e. ann.lehman-at-trincoll.edu




From: =shAf= :      mshaf-at-darkwing.uoregon.edu
Date: Thu, 29 Jun 2000 14:47:18 -0700
Subject: SEM: LaB6-to-anode distance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Hai Li:
Here is some references which claim MgAl2O4 is a
good substrate for growing perovskites:

Miura, S., Yoshitake, T., Matsubara, S., Miyasaka, Y., Shohata, N.
and Satoh, T., Epitaxial Y-Ba-Cu-O Films on Si with Intermediate
Layer by RF Magnetron Sputtering, Appl. Phys. Lett. 53:1967-1969
(1988).

Wu, X.D., Inam, A., Hegde, M.S., Wilkens, B., Chang, C.C., Hwang,
D.M., Nazar, L., Venkatesan, T., Miura, S., Matsubara, S., Miyasaka,
Y. and Shohata, N., High Critical Currents in Epitaxial YBa2Cu3O7-x
Thin Films on Silicon with Buffer Layers, Appl. Phys. Lett.
54:754-756 (1989).

My buddy, Darrell Schlom at Penn State tells me that using
perovskite substrates is much better:

Y. Jia, M.A. Zurbuchen, S. Wozniak, A.H. Carim, D.G. Schlom, L-N.
Zou, S. Briczinski, and Y. Liu, "Epitaxial Growth of Metastable
Ba2RuO4 Films with the K2NiF4 Structure," Applied Physics Letters 74
(1999) 3830-3832.

Best regards,
Mike Urbanik
www.crystalguru.com


} } Subj: information about MgAl2O4


For a tungsten emitter, my JEOL operators' manual implies to raise
the anode closer to the Wehnelt for low keV applications (less than
15keV). The manual does not mention this distance for the LaB6
emitter option. Any suggestions?

TIA and cheerios, =shAf= :o)

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - mshaf-at-darkwing.uoregon.edu
Geological Science's Electron Probe Facility - University of Oregon
http://epmalab.uoregon.edu/






From: Sarah Lundberg :      lundberg-at-nevada.edu
Date: Thu, 29 Jun 2000 15:21:34 -0700
Subject: Re: Cathodoluminescence detector

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We also have an Oxford MiniCL detector (attached to our JEOL-8900), however
the Research Instruments group of Oxford (CL and Cryo EM stuff) was just
sold to Gatan, so you may need to contact Gatan to find out about the
product. We are in the middle of some modifications to our unit (extension
of the light pipe), and JEOL informed me of the transaction. It doesn't
appear on the websites, however it was official as of 6-19-00. Hope the
info helps.
Sarah

Stafford McKnight wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Stephan,
}
} We have an Oxford MiniCL detector attached to a JEOL 6300 -
} same column as the 6400. This is an inexpensive detector which
} suits our needs but there are other suppliers.
} No matter who you obtain one from, give some thought to what
} ports are free and to your analytical working distance (presume you
} have EDS or WDS). On the 6400 this is most likely 15mm - some
} ingenuity from Oxford or other supplier will be needed to clear the
} pole-piece eg an inclined configuration. It would be worth it - rather
} than have to alter the sample height for quant analysis.
}
} Regards
}
} Stafford
} Stafford McKnight
} Geology & Metallurgy
} University of Ballarat
} ph 03 53279262
} fax 03 53279144

--
Sarah A.W. Lundberg
Electron Microanalysis and Imaging Laboratory
Department of Geoscience, UNLV
4505 S. Maryland Parkway Box 454010
Las Vegas, NV 89154-4010

EPMA Lab (702) 895-2660
SEM Lab (702) 895-2462
Office (702) 895-1134
Fax (702) 895-4064
Dept. Office (702) 895-3262

lundberg-at-nevada.edu
http://www.unlv.edu/Colleges/Sciences/Geoscience/EMP.htm

Home:
8025 W. Russell Rd. Apt. #1117
Las Vegas, NV 89113
(702) 871-9635






From: Gordon Vrololjak :      gvrdolja-at-nature.Berkeley.EDU
Date: Thu, 29 Jun 2000 15:53:56 -0700 (PDT)
Subject: Re: Printer information please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hellow Richard,
HP Laserjet 2100TN is used in our laboratory and it has a very high stated
resolution of 1200dpi. The printer gives excellent output at its highest
setting, and lower. The level of detail I obtain from the printer is
sufficient to show all the information from TEM images. However, I
believe the final output from a photograph is still slightly
better. Also, photographs have very nice glossy or matt finishes which
I haven't seen on laser printouts.

\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\
Gordon Ante Vrdoljak Electron Microscope Lab
ICQ 23243541 http://nature.berkeley.edu/~gvrdolja 26 Giannini Hall
gvrdolja-at-nature.berkeley.edu UC Berkeley
phone (510) 642-2085 Berkeley CA 94720-3330
fax (510) 643-6207 cell (510) 290-6793

On Thu, 29 Jun 2000, richard black wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I run a microscopy unit which is completely digital. I have a problem
} with printing out images. I am planning to upgrade cameras to 10 or 12
} bit megapixel units, but am already restricted by printers. Laser
} printers seem to deal with greyscales by halftoning at 100-150dpi which
} is unsatisfactory for small prints. Are there any true greyscale
} printers available, that manage better than 300dpi and better than 256
} grey levels? Is there any advantage in having more than 256 grey
} levels? Sorry if this is a bit dim, but there are so many self styled
} experts who give conflicting opinions I end up totally confused.
} Thank you.
} Richard Black
} Nottingham, England
} richardblack-at-cwcom.net
}
}
}
}





From: Pbgrover-at-aol.com
Date: Thu, 29 Jun 2000 19:55:03 EDT
Subject: re: Balzers contact info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ann,

Balzers stuff is now handled by a company called Techno Trade. If you can't
find them on the 'net or find their address, email me and I'll dig through my
files and find their address for you.

Cheers :o)

Paul Grover
Chief Microscopist and Bottle Washer
Microvista Laboratory
Lafayette, IN




From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Thu, 29 Jun 2000 17:30:47 -0700
Subject: Olympus 100X/1.4 PlanAPO available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have one new unit Olympus 100X/1.4 PlanAPO 160 tube length objective
for sale. This is a very rare objective to find.

Mine is new, unused. Perfect. List price is over $5,000 (whew).

Asking $3750. Pls decode anti-spam email or telecon me at
916.791.8191, fax at 916.791.8186.

gary g





From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Thu, 29 Jun 2000 19:23:54 -0700
Subject: Re: SEM: LaB6-to-anode distance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My Amray says the same thing. As long as the gun's Whenelt end
is the same distance for each emitter type, the spacing should
be the same. I used 4mm {=5KV, 6mm {=12KV and 8mm
for } 12KV.

gary g.


At 02:47 PM 6/29/00, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: John Foust :      jfoust-at-threedee.com
Date: Thu, 29 Jun 2000 21:57:09 -0500
Subject: What makes up a phase contrast set?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I'm selling a perhaps incomplete set of parts for a phase
contrast setup for a B&L microscope. I don't have any B&L
equipment, and it was part of a lot I got a sale....

http://cgi.ebay.com/aw-cgi/eBayISAPI.dll?ViewItem&item=369625459

What other components would've made up the entire phase contrast
set?

- John





From: Jintamas SUWANJARAT :      sjintama-at-ratree.psu.ac.th
Date: Fri, 30 Jun 2000 14:22:13 +0700 (GMT)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sir,
Please subscribe, thanks.





From: =?iso-8859-2?Q?Krzysztof_Jan_H=FCbner?= :      hubner-at-IOd.krakow.pl
Date: Fri, 30 Jun 2000 12:12:50 +0200
Subject: Conference "ADI - FOUNDRYS OFFER FOR DESIGNERS '

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


INTERNATIONAL SCIENTIFIC CONFERENCE

ADI - FOUNDRYS OFFER FOR DESIGNERS
AND USERS OF CASTINGS

23-24 NOVEMBER 2000
CRACOW – FOUNDRY RESEARCH INSTITUTE

CIRCULAR NO. 1

HOSTED BY:
Instytut Odlewnictwa
30-418 Kraków, Zakopiańska 73

CO-ORGANISER:
Ministry of Economy
Polish Foundrymen’s Technical Association
Foundry Economic Chamber
SCIENTIFIC COMMITTEE

List of persons invited to participation in the Scientific Committee of the
Conference:
Prof. Jerzy PIASKOWSKI – Poland
Prof. Czesław PODRZUCKI – Poland
Prof. Edward GUZIK – Poland
Prof. Jan RĄCZKA – Poland
Prof. Ryszard KOZŁOWSKI - Poland
Prof. Mieczysław KACZOROWSKI - Poland
Prof. Jouko.J. VUORINEN – Finland
Prof. Eduard DORAZIL – Czech Republic
Prof. Lubomir BECHNY – Slovakia
Prof. Jorge A. SIKORA - Argentina
dr Franco ZANARDI - Italy
dr Matti JOHANSON – Finland
dr Jose R. GUIRIDI - Spain
dr Konstantin UZŁOW - Ukraine

ORGANISING COMMITTEE
Dr Eng. Jerzy TYBULCZUK - Chairman
Dr Eng. Adam KOWALSKI – Secretary
Dr Eng. Józef TURZYŃSKI
M.Sc.Eng. Krystyna ŁUSZCZKIEWICZ
Eng. Marta KONIECZNA
M.A. Krystyna BANY-KOWALSKA
M.Sc.Eng. Andrzej PYTEL
Eng. Janusz CUPIAŁ

PURPOSE AND SUBJECT OF CONFERENCE
The purpose of the Conference is to enable the national and foreign research
& development centres as well
as industrial units to present their latest achievements in developing the
heat treatment technology,
metallographic examinations, and quality control systems for high-grade
ductile iron - ADI in particular.

Therefore you are encouraged to submit your proposals of papers dealing with
the following subjects :
Foundry
New grades of ductile iron, their properties and technologies of production.
Methods of spheroidising and inoculation treatments
Metals science
Electron microscopy. X-ray phase analysis. Quantitative metallography. The
techniques of colour etching
and their application in metallography. New methods and tools to examine the
structure of ductile iron and
ADI.
Heat treatment
Methods of ductile iron heat treatment. Installations. Methods of heat
treatment control. Heat treatment
parameters. Technologies of ADI fabrication.
Quality
Quality systems according to ISO and EN Standards used in production of
ductile iron and ADI. Control of
production process. Statistical quality control.

PRESENTATION OF PAPERS
Papers will be presented during the plenary meeting (time for presentation
15-30 minutes).
Language of the Conference : English and Polish

INSTRUCTIONS FOR AUTHORS OF THE PAPERS
Paper volume : 6-8 pages including tables and figures on A4 paper. Even
number of pages.
Margins : left - 2.5 cm, right - 2.5 cm, upper - 2.5 cm, lower - 2.5 cm.
Write your text in editor Word 6 or Word 97, single line spacing, font Times
New Roman 12 pt.
The starting line of a new paragraph should be indented by 1 cm..
Figures, photographs and tables should run in the text of the paper.
Captions should be typed in bold and
centred.
Equations should be centred leaving one free spacing above and below the
equation.
The pages should be numbered consecutively using soft pencil.
References : published literature cited in the text should be quoted using a
number in square brackets.
Abstract : Please start with an abstract of up to 50 words.

SAMPLE OF PAPER FORMAT

On first page (only !) leave free space from the top of 10 single line
spacings

TITLE OF PAPER IN BOLD CAPITALS (centred, 14 pt)

space 2 x 1

Name & Surname (bold, centred, 12 pt)
Affiliation, e.g. Foundry Research Institute (italics, centred, 12 pt)
Place, e.g. Kraków

space 2 x 1
ABSTRACT (bold capitals, centred, 12 pt)
space 1 x 1
Text follows - approximately 50 words, indentation of 2.5 cm on the right
and left, 12 pt.

space 2 x 1
1. INTRODUCTION (bold capitals, 12 pt)
Text justified written in single line spacing.
2. FIRST SUBTITLE (bold capitals, 12 pt)
Text justified written in single line spacing.
2.1. Second subtitle (bold, 12 pt)
Text justified written in single line spacing.
space 3 x 1
REFERENCES (bold capitals, 12 pt)
OTHER EVENTS

during the Conference, the universities, industrial plants and companies
will have an opportunity to
display their products, research methods, manufacturing techniques and
computer programmes in the
form of short (up to 15 minutes) presentations and/or exhibitions - all
those who are willing to take part in
the display are kindly requested to agree in advance with the Organising
Committee the subject and form
of display,
the organisers of the Conference also offer the possibility of publishing
the ready advertising and
information materials in the Conference Proceeding upon previous agreement
with the Organising
Committee.

Details along with the Conference programme will be circulated early in
October 2000 in CIRCULAR NO.
2.

For more information please contact the following persons:

Adam KOWALSKI tel. (+48 012) 2618-502
Krystyna BANY-KOWALSKA tel. (+48 012) 2618-591
e-mail: awkowal-at-iod.krakow.pl

Address for correspondence :

INSTYTUT ODLEWNICTWA
Centrum Informacji Naukowo-Technicznej i Ekonomicznej, Normalizacji i
Szkolenia

ul. Zakopiańska 73,
30-418 Kraków, Poland

fax: +48 (012) 260 08 70
e-mail: awkowal-at-iod.krakow.pl


SCHEDULE
completed participation forms should be sent by 31 October 2000
technical papers for presentation during the Conference should be sent by 30
September 2000
proposals of promotion during the Conference should be sent by 15 October
2000


REGISTRATION FEE
(Note : accomodation is NOT included)
full rate: 200 $
reduced rate (for participants presenting papers and posters) 150 $
The cost of promotion is to be agreed.

Please make you cheque payable to :
BPH I Oddział w Krakowie
Account No. : 10601376-320000033388

==========================================================================

International Conference
ADI - AN OFFER FOR DESIGNERS AND USERS OF CASTINGS.
23 - 24 November 2000


PARTICIPATION FORM

Family Name .......................................................

Surname ..............................................................

Title.......................... Post ..................................

Company :

Name ................................................................

Address ..............................................................

...........................................................................
Fax.......................E-mail......................................

I declare my participation in the Conference
and wish to submit a technical paper
Yes ? No ?

I wish to make promotion of my Company
Yes ? No ?

Title of paper ......................................................

..........................................................................

Author(s) ............................................................

Signature .............................................................
Please return the completed participation form to the following address:

INSTYTUT ODLEWNICTWA
Centrum Informacji Naukowo-Technicznej
i Ekonomicznej, Normalizacji i Szkolenia
ul. Zakopiańska 73, 30-418 Kraków, Poland









From: SGKCCK-at-aol.com
Date: Fri, 30 Jun 2000 04:26:25 EDT
Subject: M&M2000- Philadelphia

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Here in Philadelphia we are all looking forward to the upcoming M&M meeting
in August. The meeting promises to be the best yet. To make everyone's
arrival and traveling as pleasant as possible we recommend you visit the
Philadelphia Airport Web site at http://www.phl.org/. Here you will be able
to go directly to ground transportation and find the easiest and least
expensive ways to get from the airports to the hotels in the city. The ride
from the airport to the hotels is approximately 15-20 minutes. Depending on
the transportation you choose the prices will range from $8.00 on up.
Please feel free to E-mail me with any specific questions you may have and I
shall try to assist.
We look forward to seeing everyone in August and have a great trip.

Sincerely,

Stacie Kirsch
LAC Chair




From: Everett Ramer :      Everett.Ramer-at-netl.doe.gov
Date: Fri, 30 Jun 2000 07:34:48 -0400
Subject: Procedure for handling LN2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I will getting my first SEM/EDS shortly and to obtain the required operating permit from my safety group I need a written procedure for handling LN2. Specifically I need a written procedure for filling the 3 L dewar on the EDS detector from a 50 L dewar mounted on a cart. I plan to make the transfer by using a lab source of N2 to pressurize the 50 L dewar and have obtained all the valves and fittings required to do this from another SEM lab. I would appreciate copies of the procedure.
Thanks,

Everett Ramer
National Energy Technology Laboratory
P.O. Box 10940, Cochrans Mill Road
Pittsburgh, PA, USA 15236-0940
Voice: 412-386-4920
FAX: 412-386-4806
ramer-at-netl.doe.gov





From: JHumenansky-at-phi.com
Date: Fri, 30 Jun 2000 06:48:35 -0500
Subject: Re: SEM: LaB6-to-anode distance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



The JEOL 840 that I am familiar with (this most likely applies to most JEOL
SEM's) sets the tip whenelt distance within a narrow range and as I recall
this distance is correct when the auto bias emission current is about
100ua. This makes it possible to fine tune the first cross over with the
manual bias control so that at 20kV the emission current would be adjusted
with the manual bias control to be about 8-9 and the emission current would
be about 10=20uA. The effects of this can be seen viewing the emission
pattern. The spot size will be smaller using a high number manual bias
number. If the auto bias emission current was say only 50-60uA when the
SEM was operated at 1kV accelerating voltage it would be more difficult to
fine tune the first cross over spot size because the bias number would be
about 0-1 and the emission current would be to low. Increasing the bias
number decreases the emission current. If the emission current in auto
bias is less than 100uA the whenelt is rotated slightly CW to shorten the
distance. Manual bias is the usual way to operate the LaB6 filament while
auto bias is usually used with a W filament.

Because the LaB6 source is much brighter than W filament, the first cross
over at the source can be fine tuned with the LaB6 cathode which results in
smaller spot size on the sample ie. better resolution with sufficient
current to have good signal to noise ratio.

If you use a W filament and try to set up manual bias the same way as you
would for LaB6 you will see that the signal to noise is worse using high
manual bias number (less emission current). The LaB6 filaments are
usually operated in manual bias while the W filaments are normally used
with auto bias.

On the 840 there are either two accelerating anodes, a high anode for low
kV and a low anode for high kV which is the case if you have an elctro
static beam blanker installed or a single anode that can be raised for low
kV or lowered for high kV operation. The anode in this case snaps into the
correct position. Hope this helps.

John Humenansky/Staff Scientist
Physical Electronics, Inc. (PHI)
6509 Flying Cloud Drive
Eden Prairie, MN 55344
952-828-6387





From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Fri, 30 Jun 2000 09:04:58 -0400
Subject: Re: resin shelf life

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,
I helped empty on old lab and came across a dozen never-opened bottles of
Epon-812 (yes, the real thing). You can well imagine how long they had sat
on the shelf. My question to you is....what do I do with it? My gut
reaction is to give it to our Life Safety guys to haul away, but if its
still good that would be a waste. Short of opening one of the bottles and
making a test batch, is there any rule of thumb out there?
The bottles have a WPE of 160. They are still in the original plastic
wrap and have an EMS label. Maybe Stacie has a thouhgt on this (if she's
not overwhelmed with planning for M&M).

Thanks in advance,
Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175






From: Angela Klaus :      avklaus-at-amnh.org
Date: Fri, 30 Jun 2000 10:44:24 -0400
Subject: Re: Balzers Contact Info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Ann...

The Bal-Tec representative is TechnoTrade International. 603-622-5011.

Best,

Angela

At 02:02 PM 06/29/2000 -0400, you wrote:

} I think this query was posted recently but I haven't been able to dredge up
} the replies in the list archives - too recent, maybe.
}
} I'm looking for contact info for a colleague, for Balzers or Bal-tec. He
} inherited a B. carbon evaporator model CED 030 (also: 3U-G03-751/132). Can
} anyone help out with either contact info for the company, or a copy of the
} manual?? We will pay copy/ship charges.
}
} Thanks, much.
}
} Ann Hein Lehman
} Manager, EM Facility
} Trinity College
} Hartford CT 06106
} v. 860-297-4289
} f. 860-297-2538
} e. ann.lehman-at-trincoll.edu

---------------------------------------------
Angela V. Klaus

Director, Core Imaging Facility
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 USA

Email: avklaus-at-amnh.org
Voice: (212)769-5977
Fax: (212)496-3480
---------------------------------------------




From: Greg Erdos :      gwe-at-biotech.ufl.edu
Date: Fri, 30 Jun 2000 11:42:19 -0400
Subject: Re: resin shelf life

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have used Epon-812 that is over 25 years old with no problem. This is
valuable stuff so I hope you do not through it out. Somebody out there
will surely take it off your hands. I have enough to last me the rest of
my career

At 09:04 AM 06/30/2000 -0400, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Gregory W. Erdos, Ph.D.
Assistant Director, Biotechnology Program
PO Box 110580
University of Florida
Gainesville, FL 32611

Ph. 352-392-1295 Fax: 352-846-0251




From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Sat, 30 Mar 1996 07:57:02 +0000
Subject: Re: Procedure for handling LN2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, Everett. I guess I am glad I am not working for a government anymore.
Otherwise I would probably be having to fill out similar forms. It would be
nice if the folks that review those forms knew what the issues were. I am
always afraid that they will know just enough to be dangerous and may
actually require some unhealthy procedures. But perhaps they are sensible
and only require that you have something officially in writing to document
your best practice.

I found the following post in my files from 4 years ago. I hope Larry does
not mind me reposting it. His conclusions are quite interesting.

Warren S.



} Wil's recent comment on the safety hazards of distilled water brought to
} mind some peculiar safety regulations here in MD. In reference to liquid N
...
} officer listening in will recommend new safety procedures requiring
} protective booties!
} In the end, we can't legislate common sense, nor can we abdicate
} responsibility to those above.

Try getting your safety officer to conduct an experiment:

1. Hold out hand,
2. Pour a small volume of liquid N2 over hand
3. Now the interesting bit - put on a glove, and pour the same quantity of
liquid N2 into glove.
4. Phone for ambulance.

The point is that a brief contact causes no problems, but if the contact is
continued you get a nasty burn.

Gloves, goggles, masks (and shoes) are actually more dangerous when handling
liquid N2 than sandals and no protection. And clothes are actually more
dangerous than being naked. Get the safety officer to experiment. With a
little persuasion you can probably convince the safety officer that when
handling liquid N2, everybody should be naked.

More seriously, bureaucrats, administrators and the inexperienced should
talk to somebody who has real knowledge.

--------------------------------------------------------------
Dr. Larry Stoter
Technesis
17, Rocks Park Road, Uckfield, E. Sussex, TN22 2AT, United Kingdom
Larry-at-teknesis.demon.co.uk
--------------------------------------------------------------





From: Kristen Lennon :      kalen-at-citrus.ucr.edu
Date: Fri, 30 Jun 2000 11:47:39 -0700
Subject: immunofluorescence anti-fade

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) id NAA12520
for dist-Microscopy; Fri, 30 Jun 2000 13:53:29 -0500 (CDT)
Received: from no_more_spam.com (sparc5 [206.69.208.10])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with SMTP id NAA12512
for "MicroscopyFilteredEmail-at-msa.microscopy.com"; Fri, 30 Jun 2000 13:52:59 -0500 (CDT)
Received: from navel.ucr.edu (navel.ucr.edu [138.23.225.12])
by ultra5.microscopy.com (8.9.1b+Sun/8.9.1) with SMTP id NAA12505
for {Microscopy-at-MSA.Microscopy.Com} ; Fri, 30 Jun 2000 13:52:47 -0500 (CDT)
Message-Id: {200006301852.NAA12505-at-ultra5.microscopy.com}
Received: from batch3122 ([138.23.190.212]) by navel.ucr.edu with SMTP
for Microscopy-at-MSA.Microscopy.Com; Fri, 30 Jun 2000 11:48:07 -0700
X-Sender: kalen-at-pop.ucr.edu
X-Mailer: QUALCOMM Windows Eudora Pro Version 4.0


Hi All,
I'm asking this on behalf of a colleague, so please bear with me if I
don't have all of the details.
Does anyone out there have experience with Vectashield as an anti-fade
agent for viewing immunofluorescence? Specifically, we are looking at plant
tissue in which we are using a secondary antibody conjugated to FITC and
have found that the controls (treated only with buffer) show
"autofluorescence" after mounting with Vectashield, when they were totally
non-fluorescent when mounted in buffer. Granted, the Vectashield did expire
last month, but this is something that we've seen before with this product.
Should I advise my colleague to just take pictures as quickly as humanly
possible because we can't seem to rely on this anti-fade agent? I
appreciate any input you may have.
Have a great Independence Day weekend (for those of you in the US),
Kristen
Kristen A. Lennon
Cell, Molecular & Developmental Biology Group
Department of Botany & Plant Sciences
University of California
Riverside, CA 92521
kalen-at-citrus.ucr.edu
909-787-4525




From: McLean, Dorrance :      dmclea-at-sandia.gov
Date: Fri, 30 Jun 2000 13:21:51 -0600
Subject: Mat. Electropolishing.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Dear List,
} Does any one have the dummies guide to electropolishing?? Some years ago
} (3), I had a trouble shooting guide for electropolishing. This list had
} about 9 or 10 items, problems /solutions on it and it was very
} helpful...but it has found one of those very safe filing places (i.e. I
} can't find it!) If any one has access to this list or something similar,
} I would appreciate it if you could send me a copy.
}
} Many thanks in advance.
}
} Dorrance
} PS I haven't been keeping up with the weather reports lately but it's 95
} and clear in beautiful Livermore, California.
}




From: David Henriks :      Henriks-at-CompuServe.COM
Date: Fri, 30 Jun 2000 16:27:32 -0400
Subject: Mat. Electropolishing.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dorrance:

I do not have a "dummies" guide as such, but I do have copies of Bernie
Kestel's 66 page report "Polishing Methods for Metallic and Ceramic TEM
Specimens". The "polishing" refers to electropolishing. I would be happy
to send this out to you if you do not have it already. I also have dozens
of other papers dealing with electropolishing and hundreds of papers
dealing with specimen preparation. I can email to you a list of the other
papers if you'd like to breeze through that. Let me know.

DISCLAIMER: South Bay Technology produces equipment and supplies for
specimen prepration and, therefore, has a vested interest in promoting
their use.

Best regards-

David
Writing at 1:13:58 PM on 06/30/2000

***************************************************************************
************************

David Henriks
Vice President TEL: 800-728-2233
(toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499
San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com

***************************************************************************
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.

Message text written by "McLean, Dorrance"
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


} Dear List,
} Does any one have the dummies guide to electropolishing?? Some years ago
} (3), I had a trouble shooting guide for electropolishing. This list had
} about 9 or 10 items, problems /solutions on it and it was very
} helpful...but it has found one of those very safe filing places (i.e. I
} can't find it!) If any one has access to this list or something similar,
} I would appreciate it if you could send me a copy.
}
} Many thanks in advance.
}
} Dorrance
} PS I haven't been keeping up with the weather reports lately but it's 95
} and clear in beautiful Livermore, California.
}

{




MicroscopyListserver Archive Email Extraction Software Version NJZ07060908

Return to Microscopy Listserver Home Page