From: Gillmeister, Russ :      RGillmeister-at-crt.xerox.com
Date: Fri, 31 May 2002 09:10:21 -0400
Subject: Optical Microscopy of Fine Wire in Microelectronic Application

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
n


Peter,
We have an optical based micrometer attached to our Zeiss Axiophot and
Axiovert stage. this gives Z in 1 micron increments. The units are made by
Boeckeler Instruments who did the modification / attachment. The sensitivity
will of course depend on the quality of the stage drive.
Russ Gillmeister
Xerox
~~~~~~~~~~~~~


-----Original Message-----
} From: Peter Tomic [mailto:PTomic-at-anadigics.com]
Sent: Thursday, May 30, 2002 1:30 PM
To: Microscopy Listserver (E-mail)


Folks;

Has anyone found an optical microscopy solution to measuring microelectronic
bond wire loop height and length in one instrument? Our problem is that we
want to measure the length of a wire [.001" diameter] that has been formed
into a loop connecting two points in a ckt. This is generally not a
critical measurement but at our frequency range small variations in this
formation can mean large electrical parasitic effects.

Regards,

Peter Tomic
Anadigics, Inc.


From daemon Fri May 31 09:10:46 2002



From: Carlton, Robert :      robert.carlton-at-elan.com
Date: Fri, 31 May 2002 10:05:18 -0400
Subject: Correction: PSM Meeting (Longwood Gardens, June 6)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To all,

There is a typographical error in the meeting notice for the Philadelphia
Society of Microscopy notice for our meeting at Longwood Gardens on June
6th. The email address should read

jreffner-at-rohmhaas.com


We put too many r's in the notice. Alternatively, you can simply reply to
me at this email address.

Our apologies for the inconvenience.

Robert A. Carlton
Elan Drug Delivery, Inc.
3500 Horizon Drive
King of Prussia, PA, 19406
610-313-1360
robert.carlton-at-elan.com



********************************************************************
This communication and any files transmitted with it
may contain information that is confidential, privileged
and exempt from disclosure under applicable law.
It is intended solely for the use of the individual or entity
to which it is addressed. If you are not the intended
recipient, you are hereby notified that any use,
dissemination or copying of this communication is strictly
prohibited. If you have received this communication in
error, please notify the sender. Thank you for your co-operation.
********************************************************************



From daemon Fri May 31 09:28:47 2002



From: Greg Strout :      gstrout-at-ou.edu
Date: Fri, 31 May 2002 09:19:43 -0500
Subject: Re: Information on Moire Fringes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Have you seen.....

Reyes-Gasga J., Tehuacanero S., and Yacaman M. Jose. 1998. Moire
Patterns in High Resolution Electron Microscopy Images of MoS2.
Microsc. Res. Tech. 40:2-9, 1998.
--
==================================================================
Greg Strout
Electron Microscopist, University of Oklahoma
WWW Virtual Library for Microscopy:
http://www.ou.edu/research/electron/www-vl/
e-mail: gstrout-at-ou.edu
Opinions expressed herein are mine and not necessarily those of
the University of Oklahoma
==================================================================

Jeremy Sanderson wrote:


------------------------------------------------------------------------

The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

-----------------------------------------------------------------------.

Dear All,
Can anyone please point me in the direction of easily
understandable information on the nature of Moire
fringes. I have a good article from Scientific
American dated 1963(!), but have almost no other
information.
I have tried Google and a search, but need to sort out
the good from the bad/misleading. Any leads would be
much appreciated.
Thanks, Jeremy
jb_sanderson-at-yahoo.com

__________________________________________________
Do You Yahoo!?
Everything you'll ever need on one web page
from News and Sport to Email and Music Charts
http://uk.my.yahoo.com





From daemon Fri May 31 11:17:35 2002



From: John Twilley :      jtwilley-at-sprynet.com
Date: Fri, 31 May 2002 12:16:34 -0400
Subject: Re: legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This thread seems to reflect a common phenomenon that accompanies
technological progress: in effect, that when a more objective and
quantifiable means of accomplishing something becomes possible, it is held to
a higher standard than the technology or prior art that preceded it. For
example, DNA evidence has been subjected to a level of quantified review that
fingerprinting has generally not. Newly synthesized drugs are routinely
evaluated prior to legalization in ways that botanicals have seldom been
scrutinized.

Since digitized images are readily subject to quantifiable changes that, as
Ken says, may be described numerically, it doesn't strike us as odd to suggest
that changes made to them should be scrutinized in that way. And yet every
step in conventional photography is susceptible to substantial variations, the
sum of which is probably beyond quantifying. (Imagine! One is steeping a
sheet of film in chemical after chemical in an attempt to produce a stain that
resembles something seen by the eye!)

What of that hallowed original negative? I wonder how often photographic
evidence has been subjected to a rigorous review of how each step in its
storage and development affected its response (log e) characteristics,
contrast, density range, etc. And what about color?!. How long did time and
storage temperature act on the silver halide grain ripening? Were water
quality, time, temperature and agitation quantified at each step of
development, rinsing and fixing? What was the age of each of the baths and
the number of prior images processed at the time of this development? Was the
image tested for residual fixer or halide? How has this stained piece of
plastic and gelatin been stored since? How much highlight fading has it
undergone as a result of sulfur-containing gases and peroxides acting on the
image silver? Now, if dealing with a print instead of a negative, simply
repeat each of these variables and add them to the calculation. Next, if
there was intentional manipulation of the development in order to increase the
capabilities of the film, quantify that.

Leaving aside the enormous variations in color stability among different film
types, what of the inherent ability of dyes and pigments to render color? As
every critical photographer knows, for every film / print combination there
are certain colors that cannot be rendered accurately. Similarly, the means
by which manufacturers bias the intrinsic sensitivities of their CCD chips to
render an approximation of what the eye sees are not perfect (nor accessible
to the user, as they often form the basis for proprietary advancements in
their equipment.)

Along with prompting a review of what may be deemed acceptable within the
potentials of new imaging technology these advances should, perhaps, provoke
some thought about what the real level of accountability has been in regard to
the older technology.

John Twilley
Conservaton Scientist

Ken Converse wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Barbara,
} A number of years ago I had a customer in a state crime lab and he told
} me that any digital images had to be captured on a WORM drive (Write
} Once, Read Many), now CD-R. This was considered the rough equivalent of
} a photographic negative in court.
}
} Also, there were several threads over the past couple of years about
} enhancing (altering) data. I believe the main thrust was that you had
} to have the original data and also be able to describe (preferably with
} equations) what was done to alter the data to its final state. The
} threads go into much more detail.
}
} Ken Converse
} owner
} Quality Images
} third party SEM service
} Delta, PA
}
} Ask-A-Microscopist wrote:
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } } Status:
} } }
} } } Hi,
} } }
} } } A colleague asked an interesting question the other day: were there
} } } any legal requirements for microscope images?
} } }
} } } The only issues which I have seen are the following:
} } } 1. The standard format set by MSA is TIFF.
} } } 2. Ethically, an image can be processed for improved publication but
} } } not to any extent which corrupts data. Better microscopy is strongly
} } } preferred over processing.
} } }
} } } Do any of you know of any other legal ramifications?
} } }
} } } Thanks in advance for any input.
} } }
} } } Best regards,
} } } Barbara Foster
} } } Microscopy/Microscopy Education
} } } 125 Paridon Street, Suite 102
} } } Springfield, MA 01118
} } } PH: 413-746-6931 FX: 413-746-9311 Web: www.MicroscopyEducation.com
} } }
} }
} }





From daemon Fri May 31 14:00:40 2002



From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Fri, 31 May 2002 08:54:11 -1000 (HST)
Subject: Re: Fwd: legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, All-

The Education Committee of the Microscopy Society of America has recently
formed a new subcommittee on the ethics of digital image manipulation,
which intends to promote discussion of this issue. At M&M2002 in Quebec
City I'll be bringing it up during the Problem Solving with the Experts
session, which is scheduled for Tuesday at 8:30 am.

Currently, as far as I know, there are no legal or ethical standards for
digital image presentation except in clinical and forensic fields where
images may be used in legal cases. If anyone knows of any other standards,
please let me know.

I have some opinions and ideas, and I had planned to bring it up in this
list soon so I could get an idea how all of you felt, and then summarize
and invite more discussion at the meeting. I welcome any and all input!

I'll have a lot more to say as soon as I have time to write, but this is
as good a time as any to let the debate begin!


} } A colleague asked an interesting question the other day: were there
} } any legal requirements for microscope images?
} }
} } The only issues which I have seen are the following:
} } 1. The standard format set by MSA is TIFF.
} } 2. Ethically, an image can be processed for improved publication but
} } not to any extent which corrupts data. Better microscopy is
} } strongly preferred over processing.

Aloha,
Tina

http://www.pbrc.hawaii.edu/bemf/microangela

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************



From daemon Fri May 31 14:00:42 2002



From: max.sidorov-at-amd.com
Date: Fri, 31 May 2002 11:51:45 -0700
Subject: Information on Moire Fringes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jeremy,
I just did a google on this. Second item at the top of the search list:
http://don.sci.mu.edu/classes/L1980210.htm

{- it's about gratings but for crystals it's the same principle. Moire fringes are interference patterns produced by two overlapping lattices or gratings.

Hope this helps,

Max
__________________________________
Max Sidorov
Materials Technology Development
Advanced Micro Devices

max.sidorov-at-amd.com



-----Original Message-----
} From: Jeremy Sanderson [mailto:jb_sanderson-at-yahoo.com]
Sent: Friday, May 31, 2002 3:49 AM
To: Microscopy-at-sparc5.microscopy.com


Dear All,
Can anyone please point me in the direction of easily
understandable information on the nature of Moire
fringes. I have a good article from Scientific
American dated 1963(!), but have almost no other
information.
I have tried Google and a search, but need to sort out
the good from the bad/misleading. Any leads would be
much appreciated.
Thanks, Jeremy
jb_sanderson-at-yahoo.com

__________________________________________________
Do You Yahoo!?
Everything you'll ever need on one web page
from News and Sport to Email and Music Charts
http://uk.my.yahoo.com




From daemon Fri May 31 14:12:22 2002



From: James Roberts :      James.Roberts-at-mail.co.ventura.ca.us
Date: Fri, 31 May 2002 12:05:00 -0700
Subject: Re: Fwd: legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The real test is that the individual that is introducing the photo must be able to testify that the image "truly and accurately depicts what it is purported to represent." That is the question that we are asked when a photo is being introduced in court. If you can testify in court that it does so it will probably be introduced. Even an image taken with a wide angle or telephoto lens does not meet this criteria if being introduced to show size relative to distance, but would for some other uses.

A lot of people have developed rules they believe the courts would want, like the one stated here, but I am unaware of any standard that has been set down by the courts themselves (appellate or above in published opinions) other than the "accurately depicts" test. If your testimony were challenged on that issue it would be of great help to have an original that you could prove had not been altered in any way. Our department is going to a great deal of trouble to set up a system that provides this. But introduction of the image does not require that, at least with-in the jurisdictions I've worked in.

Jim

James L. Roberts
Supervising Forensic Scientist
Comparative Analysis Section
Ventura Co. Sheriff's Lab
800 S. Victoria Ave.
Ventura, CA. 93009

(805) 654-2308

James.Roberts-at-mail.co.ventura.ca.us


} } } "Ken Converse" {qualityimages-at-netrax.net} 05/31/02 05:44AM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Barbara,
A number of years ago I had a customer in a state crime lab and he told
me that any digital images had to be captured on a WORM drive (Write
Once, Read Many), now CD-R. This was considered the rough equivalent of
a photographic negative in court.

Also, there were several threads over the past couple of years about
enhancing (altering) data. I believe the main thrust was that you had
to have the original data and also be able to describe (preferably with
equations) what was done to alter the data to its final state. The
threads go into much more detail.

Ken Converse
owner
Quality Images
third party SEM service
Delta, PA

Ask-A-Microscopist wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} } Status:
} }
} } Hi,
} }
} } A colleague asked an interesting question the other day: were there
} } any legal requirements for microscope images?
} }
} } The only issues which I have seen are the following:
} } 1. The standard format set by MSA is TIFF.
} } 2. Ethically, an image can be processed for improved publication but
} } not to any extent which corrupts data. Better microscopy is strongly
} } preferred over processing.
} }
} } Do any of you know of any other legal ramifications?
} }
} } Thanks in advance for any input.
} }
} } Best regards,
} } Barbara Foster
} } Microscopy/Microscopy Education
} } 125 Paridon Street, Suite 102
} } Springfield, MA 01118
} } PH: 413-746-6931 FX: 413-746-9311 Web: www.MicroscopyEducation.com
} }
}
}









From daemon Fri May 31 14:40:31 2002



From: Clarkson Donna R Contr USAMRD :      donna.clarkson-at-brooks.af.mil
Date: Fri, 31 May 2002 14:33:45 -0500
Subject: TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are looking to purchase a digital camera system for our TEM and
would like some feedback from those of you who already utilize such systems.
Some questions I'd like to ask are:

* What make & model camera system do you have?
* What type, make & model printer(s) do you use, (e.g. inkjet,
dye-sublimation, silver halide)?
* Do you feel you get micrograph-quality resolution and images
utilizing that system?
* Ease of use--is it user-friendly or cumbersome?
* Ease and/or quality of service--are any of the components
serviceable by you? How quickly & easily is it to get a service rep?
* How long have you had this system?
* Would you recommend this system--why or why not?
* What, if anything, might you have done differently?

I'd greatly appreciate any responses. Thank-you in advance for your
help.

Donna R. Clarkson


From daemon Fri May 31 15:44:37 2002



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Fri, 31 May 2002 16:30:43 -0500
Subject: Manipulation of images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

James Roberts wrote:
=====================================================
The real test is that the individual that is introducing the photo must be
able to testify that the image "truly and accurately depicts what it is
purported to represent." That is the question that we are asked when a
photo is being introduced in court. If you can testify in court that it does
so it will probably be introduced. Even an image taken with a wide angle or
telephoto lens does not meet this criteria if being introduced to show size
relative to distance, but would for some other uses.

A lot of people have developed rules they believe the courts would want,
like the one stated here, but I am unaware of any standard that has been set
down by the courts themselves (appellate or above in published opinions)
other than the "accurately depicts" test. If your testimony were challenged
on that issue it would be of great help to have an original that you could
prove had not been altered in any way. Our department is going to a great
deal of trouble to set up a system that provides this. But introduction of
the image does not require that, at least with-in the jurisdictions I've
worked in.
==========================================================
One must always remember that the expert giving such testimony is rendering
professional judgements and opinions and if one or another party to a
dispute should believe or try to construe that the independent expert made a
professional error, which resulted in economic loss to them, then that
expert can be sued for professional malpractice just as a physician or any
other professional can be sued in our courts by anyone who feels they have
been harmed by the professional negligence of the expert.

This reality becomes a powerful force to make sure sure that any image
enhancement or alteration is done in ways that do not distort the outcome of
one's conclusions. Remember,just as for physicians and lawyers, one does
not really have to have made a profesisonal error, since someone needs only
to construe that an error was made, for them to be on the receiving end of a
law suit.

We all know that lawyers and physicians pay a great deal of money for
insurance to fund such losses. But people rendering opinions on microscopy
results tend to forget about the need for such insurance coverage, even
though many times they put their own personal assets at risk when they
render such testimony (without professional liablity insurance coverage).

Do laboratories get sued? You bet. Anyone doing this type of work should
have some kind of a program of loss prevention and risk analysis in order to
reduce the risk that any of their laboratory's results could be
misinterpreted or misconstrued by others.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================


From daemon Fri May 31 16:02:41 2002



From: Heather A Owen :      owenha-at-csd.uwm.edu
Date: Fri, 31 May 2002 15:55:04 -0500 (CDT)
Subject: SEM specimen holder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Everyone,

We are using an Hitachi S-570 SEM for examination of rock thin sections
mounted on 1" x 3" glass microscope slides. Up until now, we have been
securing the slides to a large circular (50 mm) specimen holder with
copper and/or carbon tape. We would like a holder that would not require
the use of tape, carbon paint, etc. Does anyone know if there are any
commercially available holders (similar, for example, to that found in a
Cameca Microprobe, where a metal frame secures and grounds the slide to
the base) that would work in our microscope?

Any design ideas would also be appreciated since our machine shop will
fabricate what we need as long as nothing is commercially available.

Thanks for any help,

Heather Owen

Dr. Heather A. Owen, Director
Electron Microscope Laboratory
Department of Biological Sciences
University of Wisconsin - Milwaukee
Lapham Hall, P.O. Box 413
Milwaukee, WI 53210
USA

Phone: (414)229-6816



From daemon Fri May 31 17:29:54 2002



From: Hiromi Konishi :      hkonishi-at-asu.edu
Date: Fri, 31 May 2002 15:21:34 -0700
Subject: Cerius Program

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


{fontfamily} {param} Times_New_Roman {/param} Cerius users: 5/31/02

{/fontfamily} {fontfamily} {param} Times {/param}

I am having troubles with image simulations along low-index
orientations using Cerius program (Cerius 2 Ver.4).


I redefined unit-cell and changed the orientation that I want to
calculate to a primary axis. I divided the new cell into several slices
and I calculated projected potential for each slice. However, the
resulting ED patterns have extra spots that dose not match to the
reciprocal lattice points of the original unit-cell. I thought the
extra spots are from high order Laue zone, but the spacing does not
match to that I expect in some cases.


What I would like to ask are:

(1) How can I interpret the calculated ED patterns, assuming that
Cerius does not go wrong?

(2) If Cerius code has troubles, at what conditions does it go wrong?

(3) Is there any other programs that allow us to calculate low-index
orientation images?


The msi files of the structures, log files, and the calculated
diffraction patterns, and copies of comments I received from a user and
the developer are available upon request.


I would appreciate any comments you might have.


Hiromi konishi hkonishi-at-asu.edu {smaller}

{/smaller} {/fontfamily}




From daemon Fri May 31 18:10:45 2002



From: gary.m.brown-at-exxonmobil.com
Date: Fri, 31 May 2002 18:03:15 -0500
Subject: Re: Fwd: legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


The real test is that the individual that is introducing the photo must be
able to testify that the image "truly and accurately depicts what it is
purported to represent." That is the question that we are asked when a
photo is being introduced in court. If you can testify in court that it
does so it will probably be introduced. Even an image taken with a wide
angle or telephoto lens does not meet this criteria if being introduced to
show size relative to distance, but would for some other uses.

A lot of people have developed rules they believe the courts would want,
like the one stated here, but I am unaware of any standard that has been
set down by the courts themselves (appellate or above in published
opinions) other than the "accurately depicts" test. If your testimony were
challenged on that issue it would be of great help to have an original that
you could prove had not been altered in any way. Our department is going to
a great deal of trouble to set up a system that provides this. But
introduction of the image does not require that, at least with-in the
jurisdictions I've worked in.

Jim

James L. Roberts
Supervising Forensic Scientist
Comparative Analysis Section
Ventura Co. Sheriff's Lab
800 S. Victoria Ave.
Ventura, CA. 93009

(805) 654-2308

James.Roberts-at-mail.co.ventura.ca.us


} } } "Ken Converse" {qualityimages-at-netrax.net} 05/31/02 05:44AM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Barbara,
A number of years ago I had a customer in a state crime lab and he told
me that any digital images had to be captured on a WORM drive (Write
Once, Read Many), now CD-R. This was considered the rough equivalent of
a photographic negative in court.

Also, there were several threads over the past couple of years about
enhancing (altering) data. I believe the main thrust was that you had
to have the original data and also be able to describe (preferably with
equations) what was done to alter the data to its final state. The
threads go into much more detail.

Ken Converse
owner
Quality Images
third party SEM service
Delta, PA

Ask-A-Microscopist wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} } Status:
} }
} } Hi,
} }
} } A colleague asked an interesting question the other day: were there
} } any legal requirements for microscope images?
} }
} } The only issues which I have seen are the following:
} } 1. The standard format set by MSA is TIFF.
} } 2. Ethically, an image can be processed for improved publication but
} } not to any extent which corrupts data. Better microscopy is strongly
} } preferred over processing.
} }
} } Do any of you know of any other legal ramifications?
} }
} } Thanks in advance for any input.
} }
} } Best regards,
} } Barbara Foster
} } Microscopy/Microscopy Education
} } 125 Paridon Street, Suite 102
} } Springfield, MA 01118
} } PH: 413-746-6931 FX: 413-746-9311 Web: www.MicroscopyEducation.com
} }
}
}












From daemon Fri May 31 18:14:29 2002



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Fri, 31 May 2002 18:05:15 -0500
Subject: Re: legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Actually, in addition to guidelines for generation of images,
there can be issues of data ownership and legal ramifications based
on interpretation of micrographs by the microscopists. These and
other related questions will be discussed at M&M2002 during the
Technologist's Forum Roundtable. The topic for this session is:
Legal and Ethical Issues of Data Ownership".

The panel will consist of: Bertha M Knoppers, an internationally
recognized expert in ethics and the law, as well as representatives
from both academia and industry.

Hope many of you can attend because responses to questions related
to this topic are not always as obvious as one would expect.
Debby


Debby Sherman Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
West Lafayette, IN 47907


On Thursday, May 30, 2002 8:56 PM, Ask-A-Microscopist
{zaluzec-at-sparc5.microscopy.com} wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


From daemon Fri May 31 18:14:48 2002



From: David Hall :      hall-at-aecom.yu.edu
Date: Fri, 31 May 2002 18:05:47 -0500
Subject: Re: Fwd: legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Barbara -

The National Forensic Science Technology Center maintains a site for The
Scientific Working Group for Imaging Technologies at
http://for-swg.org/it_files/swgit_guidelines.html. The site includes draft
guidelines for use of imaging technologies by criminal justice
professionals.

James Martin
Orion Analytical, LLC
www.orionanalytical.com
martin-at-orionanalytical.com


----- Original Message -----
} From: "Ask-A-Microscopist" {zaluzec-at-sparc5.microscopy.com}
To: {Microscopy-at-sparc5.microscopy.com}
Sent: Thursday, May 30, 2002 9:56 PM


There are two more significant legal aspects that might be brought up.

1. To whom does the microscope image belong as intellectual
property, and under what conditions can the image be shared between
groups, placed on a website, or published in someone's manuscript.
How does one adequately attribute the original source of the image? I
certainly have experienced occasional surprise when an old workprint
of mine has been scanned to create a new published illustration.

2. For a previously published microscope image, how much
manipulation transforms the data into a "new" image which can be
published without violating the copyright on the original publication?

Obviously we need a lawyer or two on this list. Does anyone know of
a good article or book that covers this ground?
--
David H. Hall, Ph.D.
Center for C. elegans Anatomy
Department of Neuroscience
Albert Einstein College of Medicine
1410 Pelham Parkway
Bronx, NY 10461

www.wormatlas.org
www.aecom.yu.edu/wormem

phone 718 430-2195
fax 718 430-8821


From daemon Fri May 31 22:45:25 2002



From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Fri, 31 May 2002 23:30:58 -0400 (EDT)
Subject: Re: Manipulation of images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The most basic rule is that the controls and experimentals must be
treated the same before they are imaged and that they are imaged by the
same parameters.
Following the simple rules of doing valid controls and
taking the raw data in a uniform manner is far more critical than
the post-collection manipulation.

And the problems with the image manipulation are not that images are
altered intentionally, but that imaging amateurs (who may be the most
brilliant
geneticists or endocrinologists), who don't understand digital
representations, misuse Photoshop. A very common example is the misuse of
the Auto button in the Levels or Curves menu when rescaling from 12 bits
to 8 bits or to-byte-as-shown in any of the imaging packages. And then
the question we get is how to fix the problem, not to
compound it. "Why do the WT and KO look the same on the computer when I
swear the WT looked at least 2 times brighter through the eyepiece and the
Westerns show a 10X reduction in expression in the KO?"

People who want to cheat will always find a way to do so. The real
problem is educating to avoid simple imaging errors.


-Michael

___________________________________
WORK: http://www.aecom.yu.edu/aif/




From daemon Sat Jun 1 06:33:41 2002



From: Bill Brady :      wmbrady-at-olg.com
Date: Sat, 1 Jun 2002 07:22:42 -0400
Subject: Re: [Microscopes] basic stains and chemicals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rooting Wrote:

} For Formaldehyde/Formalin, are there any replacements which are
} non-flammable?

I use "Quick Cure" for aquarium use. Works fine in most cases. It's main
ingredient is Formalin.
}
} How good is frutose for a mounting solution?

See:
http://www.microscopy-uk.org.uk/larry/sugar0.html

Wm. "Bill" Brady, Harwood MD 38°51'30"N 76°41'00"W - Its in the darkest
hour that the most stars come out.



From daemon Sat Jun 1 14:11:02 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Sat, 1 Jun 2002 13:08:58 -0600
Subject: Re: Manipulation of images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Manipulation of images is an issue that not only comes up in legal
proceedings, but in many other settings as well. One area is Pharmaceutical
research. There are very strict standards that the FDA sets regarding data
manipulation. Lab books have to be kept in certain ways to ensure that any
manipulation of data (legitimate or not) is traceable.
A few years ago, the FDA published a document (CFR21 rule 11), which
describes how a digital equivalent of a lab book has to be kept.
Unfortunately this document mixes applications, Operating systems, hardware
and SOPs, so it is very hard to implement (ask me off-line how we do that).
The essence is, that a document has to be defined as an "original" at some
point, as close as possible to the origin of the data. For a digital image,
this would probably be the time it is transferred to the computer and
displayed for the first time. ANY change to the data afterwards needs to be
documented and signed. For example, if you run a filter on the image, it
must have an audit trail and you must be able to go back to the original
image. That's a tall order for images, but it can be done. This kind of
audit trail would probably also stand in legal proceedings.

I don't think, that even this is a 100% insurance against malicious intend.
After all, you could theoretically change something in the camera
electronics, which would change the image before it was declared an original
document, but it comes pretty close.

mike

} } } } } } } WE HAVE MOVED { { { { { { { { {
please make a note of the new address below

Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Ave #300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Michael Cammer [mailto:cammer-at-aecom.yu.edu]
Sent: Friday, May 31, 2002 9:31 PM
To: Garber, Charles A.
Cc: MICROSCOPY BB


The most basic rule is that the controls and experimentals must be
treated the same before they are imaged and that they are imaged by the
same parameters.
Following the simple rules of doing valid controls and
taking the raw data in a uniform manner is far more critical than
the post-collection manipulation.

And the problems with the image manipulation are not that images are
altered intentionally, but that imaging amateurs (who may be the most
brilliant
geneticists or endocrinologists), who don't understand digital
representations, misuse Photoshop. A very common example is the misuse of
the Auto button in the Levels or Curves menu when rescaling from 12 bits
to 8 bits or to-byte-as-shown in any of the imaging packages. And then
the question we get is how to fix the problem, not to
compound it. "Why do the WT and KO look the same on the computer when I
swear the WT looked at least 2 times brighter through the eyepiece and the
Westerns show a 10X reduction in expression in the KO?"

People who want to cheat will always find a way to do so. The real
problem is educating to avoid simple imaging errors.


-Michael

___________________________________
WORK: http://www.aecom.yu.edu/aif/




From daemon Sun Jun 2 10:09:45 2002



From: Jackie :      ujtxh-at-eszett.de
Date: Sun, 02 Jun 2002 07:49:55 +0900
Subject: Great shape for summer,get there now

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


As seen on NBC, CBS, CNN, and even Oprah! The health
discovery that actually reverses aging while burning fat,
without dieting or exercise! This proven discovery has even
been reported on by the New England Journal of Medicine.
Forget aging and dieting forever! And it's Guaranteed!

Click here:
http://66.231.133.205/hgh/index.html

Would you like to lose weight while you sleep!
No dieting!
No hunger pains!
No Cravings!
No strenuous exercise!
Change your life forever!

100% GUARANTEED!

1.Body Fat Loss 82% improvement.
2.Wrinkle Reduction 61% improvement.
3.Energy Level 84% improvement.
4.Muscle Strength 88% improvement.
5.Sexual Potency 75% improvement.
6.Emotional Stability 67% improvement.
7.Memory 62% improvement.

***********************************************************

You are receiving this email as a subscriber
to the Opt-In America Mailing List.
To unsubscribe from future offers,
just click here:
mailto:affiliateoptout-at-btamail.net.cn?Subject=off




vrye


From daemon Mon Jun 3 08:22:19 2002



From: William Oxberry :      William_Oxberry-at-downstate.edu
Date: Mon, 3 Jun 2002 09:08:16 -0400
Subject: Re:TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Donna,

I recently looked into getting a digital camera for my Zeiss TEM. At a
cost of $40k-$50k I decided to stick with 4x3 1/4 film and D-19 development
(archival negs) and digitize with a good scanner (Epson perfection 2450
about $400). I can get decent prints on a laser printer or my new Epson
C80 bj printer. This printer costs $150 and is amazing. If you need faster
output printing you can buy a dedicated computer, a couple of these
scanners and several printers and still spend less than $5k and have enough
paper for a couple years. At a savings of $35k my dept chair was happy

Bill




From daemon Mon Jun 3 08:32:17 2002



From: James Hayden :      jhayden-at-wistar.upenn.edu
Date: Mon, 3 Jun 2002 09:26:23 -0400
Subject: Ethics of Digital Manipulation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This seems as good a time as any to put in my $.02.

Unfortunately, my answer would go for pages. I wrote an article on the
subject a few years back aimed at photographers and graphic artists that
may or may not be aware of the issues, especially in regard to
scientific images. It was published in the Journal of Biocommunications
and later, as an update in the Network journal of the Australian
Institute of Medical and Biological Illustration. This second paper is
reproduced (with permission) on my web site at:

http://www.biographics.org/pages/ethics.html

The paper was designed to make people aware of the issue and has a few
examples of what I termed "acceptable" and "unacceptable" manipulations,
along with information from a German case of particularly blatant
manipulation of images.

I just began a new position at The Wistar Institute in Philadelphia and
part of my mandate is to introduce the topic into required ethical
training here. This is a topic that will be with us for a long time. Any
comments on-line or off would be greatly appreciated. This thread is
certainly a good start.

Jamie Hayden

*********************************
James E. Hayden, RBP, FBCA
Manager: Microscopy Core Facility
The Wistar Institute
Room B-78
3601 Spruce Street
Philadelphia, PA 19104

office phone: (215)898-3887
cell phone: (215)514-4223
email: jhayden-at-wistar.upenn.edu



From daemon Mon Jun 3 08:58:53 2002



From: sjb27-at-cornell.edu
Date: Mon, 3 Jun 2002 09:52:29 -0400 (EDT)
Subject: LM Need cutting wheel specs for LKB Histo Knifemaker 2078

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Recently received a used LKB Histo Knifemaker 2078
(makes Ralph glass knives, not triangular glass knives)
but need to replace the cutting/scoring wheel. The
manual does not list the specifications of the wheel,
and the company that made the Knifemaker is no longer
in business making knifemakers.

Does anyone know the specs, or know a supplier for the
cutting wheels?

Thanks,
Sandra Borgardt
L. H. Bailey Hortorium
462 Mann Library
Cornell University
Ithaca, NY 14853


From daemon Mon Jun 3 08:59:36 2002



From: mail.thelinks.com :      sstouden-at-thelinks.com
Date: Mon, 03 Jun 2002 08:57:57 -0500
Subject: Re: Fwd: legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



It has been my experience in expert testimony, that things like images need
expert testimony.
Have you seen the image that is on this media(cross examiner holds media
and image in hand)? Did you take the image? Is this the image you made? Are
you sure? How many other images did you make in the course of doing this
work? What happened to them? Did you modify the image in any way? IF so,
what were the nature and extent of those modifications? Are those types of
modifications normal practice in your industry, in your lab, for or by you?
Did you do the work yourself or was a technican involved? If technican,
what is his/her name? Without the modifications, what would have been the
result? At any time during the time you started the work and ended the
work was the "object to be investigated" under your direct control? If
not, please explain.
Are the images reproducible at any time by anyone with your level of training?
What type of equipment did you use? What is the best equipment that could
have been used to make this image? What did you not use that? Were you
limited by availability in your lab? If so, would that have affected the
result?
Ok, now in your opinion what does the image labeled "such and such" on
the "media" you provided the court, the one I am holding up for the jury to
see now mean to you? ... your answer.... Is your answer subject to any
kind of doubt or question, if so, what is the nature of your doubt or
concerns? Did you use any assumptions in expressing your findings about
this image? If so, what were they?
In your opinion as a professional the image shows " that the bullet that
killed the Deer was fired from this weapon [holding it up]...at such and
such a time on such and such a date?" Is that correct?

(be careful, you can only testify that the bullet was fired from the weapon
on a date and time, not that the bullet killed the deer unless you made
that determination) Thank you.


At 08:44 AM 5/31/2002 -0400, Ken Converse wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Mon Jun 3 09:25:58 2002



From: Brad Storey :      storey-at-lanl.gov
Date: Mon, 03 Jun 2002 08:19:10 -0600
Subject: EM Position at Los Alamos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Los Alamos National Lab has an opening for a electron microscopist to work
in the Nuclear Materials Science group (NMT-16) on Pu alloys and other
actinides.

Applicants must apply directly through the official LANL web site, but they
may contact Brad Storey at storey-at-lanl.gov or 505-667-0458 with questions.
Note that this position requires US citizenship.

Summary:
The Nuclear Materials Science Group, NMT-16, of the Nuclear Materials
Technology Division is seeking a highly motivated materials scientist to
study plutonium and other actinide materials. The successful candidate
will participate in a team environment in ongoing experiments to support
Pit Surveillance, Pit Manufacturing, and Pit Certification Programs. The
primary experimental focus will be in fielding and advancing
state-of-the-art materials electron microscopy to investigate production
problems, component failures, and aging stockpile issues. They will have
access to three newly acquired instruments: a JEOL 6700F field-emission gun
SEM (WDS and EDS); a JEOL JAMP-7830 field-emission gun Auger Microprobe
(Auger, XPS, cold fracture, heating, orientation imaging microscopy, gas
reaction, AFM/STM); and a Kratos Axis-Ultra imaging XPS system (XPS, Auger,
heating/cooling, gas reaction). These instruments will be installed in the
Plutonium Facility at TA-55 in the spring of 2003, and enhance the
experimental capabilities of a well-equipped material science laboratory
that includes: a variable pressure SEM (EDS, WDS, OIM, CL spectroscopy,
hot/cold stage); a 5 spectrometer JEOL 8200 Electron Microprobe; diverse
x-ray diffraction capabilities; hardness testers; several modern optical
microscopes with digital cameras; and a nicely equipped metallography
line. The successful candidate will perform hands-on work with plutonium
and other actinides in a glovebox environment and mentor technicians in
materials science topics related to particular experiments.

Required Skills:
Extensive experience in materials characterization studies utilizing
electron-beam instrumentation and associated detectors, e.g., SEM,
microprobe, EDS, WDS, CL, OIM, etc. Recent peer-reviewed publications that
include SEM and microprobe as characterization tools. Strong knowledge of
quantification procedures relating to these techniques. Experience
maintaining state-of-the-art electron microscopy equipment and teaching
technicians and staff to operate it. Strong knowledge base in materials
science as evidenced by an advanced degree in the field and a
publication/presentation record. Demonstration of organizational skills and
good communication skills. Ability to obtain a Q clearance, which usually
requires US citizenship.

Desired Skills:
Experience in the design, production, or evaluation of nuclear weapon
systems and components. Ability to evaluate microstructures as they relate
to materials properties and processing history. Experience working with
nuclear materials in glove boxes, classified document handling at all
classification levels, and familiarity with quality assurance principles.
Experience writing successful proposals to secure funding. Active DOE
Q-clearance and PSAP approval.

Education:
MS (Ph.D. preferred) in materials science, nuclear, or metallurgical
engineering, or equivalent combination of education and experience.

Additional Requirements:
This position is subject to the requirements of the Personnel Security
Assurance Program (PSAP). All candidates invited for an interview must
consent to be in the PSAP program at the time of the interview. Only the
selected candidate will be subject to the requirements of the PSAP program,
which includes a pre-employment screening check, medical examination, and
drug test.




***************************************************
Brad Storey, Ph.D.
Team Leader, Microstructure & Microanalysis
Nuclear Materials Science Group (NMT-16)
Nuclear Materials Technology Division
Los Alamos National Lab
PO Box 1663, MS E574
Los Alamos, NM 87545

Tel.: 505-667-0458
FAX: 505-665-7815
pager: 505-996-3129
storey-at-lanl.gov
***************************************************




From daemon Mon Jun 3 09:49:22 2002



From: Greg Strout :      gstrout-at-ou.edu
Date: Mon, 03 Jun 2002 09:42:41 -0500
Subject: Ultracut Fluorescent bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know what the specs are for replacement fluorescent bulbs on
a Reichert Ultracut microtome? The specs are not in the manual and our
bulbs are so old that the printing is illegible. TIA.

--
==================================================================
Greg Strout
Electron Microscopist, University of Oklahoma
WWW Virtual Library for Microscopy:
http://www.ou.edu/research/electron/www-vl/
e-mail: gstrout-at-ou.edu
Opinions expressed herein are mine and not necessarily those of
the University of Oklahoma
==================================================================




From daemon Mon Jun 3 11:24:26 2002



From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Mon, 03 Jun 2002 09:11:44 -0700
Subject: Re: SEM specimen holder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Heather,
I also use an S-570 for lots of different kinds of samples and I have got my
shop to make all sorts of clamps, jigs and pin-stub holders over the years.
I would think that a shallow groove to match the glass slide width could be
milled in the flat holder to just fit the slide. Either that, or a vise with
a fixed side and a sliding or spring-mounted side to close on the slide.
Just make sure it is made all of non-magnetic materials. Whatever you get
them to make just needs a 4M threaded hole in the bottom to take the stage
insert that screws into the bottom of your existing specimen holder.
At 03:55 PM 05/31/2002 -0500, you wrote:
} Hi Everyone,
}
} We are using an Hitachi S-570 SEM for examination of rock thin sections
} mounted on 1" x 3" glass microscope slides. Up until now, we have been
} securing the slides to a large circular (50 mm) specimen holder with
} copper and/or carbon tape. We would like a holder that would not require
} the use of tape, carbon paint, etc. Does anyone know if there are any
} commercially available holders (similar, for example, to that found in a
} Cameca Microprobe, where a metal frame secures and grounds the slide to
} the base) that would work in our microscope?
}
} Any design ideas would also be appreciated since our machine shop will
} fabricate what we need as long as nothing is commercially available.
}
} Thanks for any help,
}
} Heather Owen
}
} Dr. Heather A. Owen, Director
} Electron Microscope Laboratory
} Department of Biological Sciences
} University of Wisconsin - Milwaukee
} Lapham Hall, P.O. Box 413
} Milwaukee, WI 53210
} USA
}
} Phone: (414)229-6816
}
Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca



From daemon Mon Jun 3 11:30:37 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Mon, 3 Jun 2002 12:24:22 -0400
Subject: RE: AO & TB O Metachromasia and Science - Personal Reflections on

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sergey,

In 1965, in a lecture by one of the virologists who worked on Flu
vaccines at Merck, Sharpe and Dhome(?) (West Point, PA), he mentioned that
early on before the discovery of the RNA nucleoids of certain viruses, AO
became briefly popular, because it differentiated between DNA and RNA
(strandedness not then an issue). For a while, green fluorescence in AO
stained cultures of DNA viruses was used to follow processes leading to
assembly and burst (CPE). Then, the axe fell. RNA viruses fluoresced green
during development. Very mixed up cytoplasm of infected cells since both
single and double-stranded RNA(?) appeared to be present. This was reported
to have been announced at a meeting (?) after a paper in which AO
fluorescence was part of the presentation. Why the problem? Well, one of
the more difficult problems in the study of any virus was getting enough
pure virus to work on. Most microbiologists then were NOT biochemists. The
use of AO in THAT context was ended. We had one of the newer Sorvall
Ultracentrifuges. Three people were permitted to use it, and most didn't
understand the principles of its operation!

Viruses were still "filterable agents".

Re: Foster R Jr, Metcalf D, Kirchmyer R., "Induction of
bone marrow colony-stimulating activity by a filterable agent in leukemic
and normal mouse serum". J Exp Med. 1968 May 1;127(5):853-66.

and Re: Cook MK.,"Cultivation of a filterable agent
associated with Marek's disease", J Natl Cancer Inst. 1969 Jul;43(1):203-12.


One key to understand is that in 1965, as far as we knew, there was
only one course in virology at the 300 level in the country. The book that
was used was a brief text that concentrated on phages and E. coli and
Hershey and Chase and, of course, the Warring Blender - at that moment in
time, also a very modern culinary machine, which my Mother wanted but had
not yet received for Xmas. Further, most undergraduate cell biology books
still hadn't incorporated Watson and Crick, except as recent historical
notes. Full incorporation took almost 20 years - ~1975.

Acta Microbiol Acad Sci Hung 1966;13(2):185-7, "Acridine orange
fluorescence of tissue cultures infected with Aujeszky's disease virus",
Bodon L, Greczi E.

The above reference was typical of reports of the use of AO in
virology. Nucleic acids, cell culture.

Of the great sexually-oriented processes then known: prophase,
leptonema, zygonema, pachynema, diplonema, diakinesis, metaphase, anaphase,
replication, transcription, translation, and osculation, only the last MIGHT
have been found on the S.A.T. It is problematic whether any of the last
four would have appeared on the graduate record exams of the time [I can't
remember!!!].

What might be called second-tier microbiologists around the world
were trying to learn the new jargon and techniques while being inundated
with NEWER jargon and techniques. Delbruck and Lauria were pushing the
envelope [and we couldn't keep track!], and they were applying the new
Watson and Crick dogma to microbiology and bacterial viruses, while others
were just beginning to investigate the 'inward workings' of the few animal
viruses that were known. The same virologist mentioned above, Richard
Malsberger at Lehigh, would demonstrate during the afternoon lab the
fundamental of what was known about Herpes simplex by initiating a full
eruption of a so-called "cold sore" on his lower lip by eating a piece of
DARK chocolate at the beginning of the 8:00am lecture the same morning [my
timing may be off after all these years!]. Question: would he do that
demo in his Immunology class today or would it still remain in Virology?

In five years at Lehigh, 3 or 4 graduate students took the virology
lab. We learned to do I.D. and L.D. 50's on mice using adapted Flu virus.
We learned hemagglutinin tests/assays, complement fixation, viral
titrations, 2-fold, 5-fold, 10-fold dilutions - things only graduate
students understood. We learned about Landsteiner, and began to feel that
only Physical chemists would ever do anything or understand anything. We
learned how to fail at cell culture using the wrong "distilled" water.
Most labs had water stills hanging on the walls in the prep rooms whose
boiler and copper cooling coils were plated with chromium. Deionizers were
modified/upgraded water softeners. If you cultured animal cells, you
cultured HeLa cells. We did CPE's and plaque assays. Those who taught and
learned were probably a decade behind those who were investigating and
reporting - perhaps little different from today.

A Leitz Ortholux with fluorite objectives, darkfield condenser and a
mercury arc illuminator with exciter and barrier filters, AND the Orthomat
fully automatic 35mm camera system with TWO film cassettes [one for color
(for fluorescence) and one for B&W (for bright field)] cost around
$6,700.00. This system was SOooo advanced that the senior graduate student
in virology had no experience with microscopy and misinterpreted the
instructions for oil immersion to suggest that the objective was to be
FILLED with oil (which he did!). The objective was returned to Germany for
cleaning. We made out own fluorescent antibody, first with Fluorescein and
then with FITC. The clonal selection theory (Burnet and others) was
something in a book that only graduate students in virology read. When I,
an anatomy student, was seen reading THAT book, it was observed that I was
wasting my time. When I took microbial biochemistry, I was considered wise
by some and as a reactionary by others.

RCA made electron microscopes and they were NOT found
everywhere! The ultramicrotome of the day was the MT-1, and the NEW, magic,
motorized MT-2! We had just received the just released those magic
balances by Mettler. Oh, to weigh a gram in a minute to 2 places of
precision! [I now have a double pan balance of my youth as "...the largest
brick-a-brack in our house".]

World War II had been over for 20 years and the transistor had just
found its way into affordable, portable radios - AM AND FM! And then, of
course there was SPUTNIK and Gagerin and the Moon! Electrophoresis was on
purified cellulose paper! Total nitrogen by Kjeldahl! Amphoteric amino
acids by acid titration, deflections, and chart recording. A Frieden
mechanical calculator with twenty columns of numbers "clunkety, clunkety,
clunkety" - - off the lab bench onto the floor - - - "clunkety, clunkety,
clunkety" - - - all through the lunch hour - - "clunkety, clunkety" - - -
lying on its side, on the laboratory floor. A magic computer in the
basement [all of it!] of the Engineering College". Stacks of cards. Boxes
of stacks.

While there were some during this time who preferred being stoned
[they get all the publicity now!], most were sober and trying to soak it all
up and get on for the ride. I grew up 3 miles from Bell Labs in Berkely
Heights (Murray Hill), NJ and 15 miles from Princeton and saw Prof. Albert
walking on the street one day as our 54 Chevy passed through town and met
two of the transistor guys at a party in Summit, NJ sometime later. We were
taught in the 50's that only three or four people understood what Prof.
Albert presented in his benchmark papers in the 20's. Relativity had more
social meaning, even then, than physical.

In any case, back to AO. As I note the history of virology now, the
use of AO to study virus infections then is not even given the space of a
footnote in current histories. Not an undeserved relegation.

Sorry for the additional ramble,

Fred

} ----------
} From: Sergey Ryazantsev
} Sent: Thursday, May 30, 2002 7:10 PM
} To: Monson, Frederick C.; Microscopy-at-sparc5.microscopy.com
} Subject: Re: Toluidine Blue O - Metachromasia and Science
}
} Fred
}
} I don't understand your point about Acridine Orange (AO)? How discovery of
}
} the RNA-viruses affects the staining and our knowledge about this
} particular staining agent? As I remember, AO staining is based on polar
} interactions, mostly with "nucleic acids". The 'color' of fluorescence
} depends from the energy transfer efficiency between donor (AO) and
} acceptor
} (which may be, may be not nucleic acid). I expect, the 'color' from AO
} stained double strands RNA would be similar to DNA. Another point here,
} as
} far as I know, RNA viruses in most cases contain "double-stranded RNA
} segments", so it's a mixture single/double-stranded RNA. I could not
} predict how it will fluorescence with AO. Nevertheless, such viruses are
} very small and I don't believe you could even see their fluorescence with
} AO on the autofluorescence background of the typical cell. The bottom
} line
} here: the discovery of the new "RNA-viruses" is not affected our knowledge
}
} of basic properties of AO and AO would work (I believe) at the same manner
}
} with this 'double-stranded RNA' as with other stuff. I don't see big
} difference between single/double-stranded-RNA and DNA. It's well known,
} that most RNAs tends to form secondary structure (call it double-stranded
} RNA fragments) under the physiological conditions. Moreover, my personal
} experience shows that it's extremely difficult to make RNA completely
} 'single-stranded': even at the presence of 8MUrea+formamide,
}
} Sergey
}
}
}
} At 09:56 AM 5/30/02, you wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Morning All,
} }
} } For those among you who are wise, you should probably
} immediately
} } jump or GOTO END!
} }
} } This is difficult to put into words without chancing a gross
} error
} } or causing some offence, but here goes anyway.
} }
} } When we "fix" a specimen taken from a biological source, we
} NEVER
} } quite know what we are doing. The entire process is 'outcome-based'
} } (Ugh!!!). How does it look? Does the fixative 'coagulate' or
} 'cross-link'
} } the myriad of chemical entities within the sample? What is the effect of
} } specimen size? What about temperature? How long to kill? How long to
} } achieve minimum preservation? Preservation of what? What will
} constitute
} } comparable samples that permit statistical comparison of specimens of
} mouse
} } and elephant liver? We preserve or fix or kill or maintain color or get
} rid
} } of it. Behind every act there lies a reason, and in consequence of every
} } act there is produced an uncountable number of variables which only
} } REPRESENT what existED in the original, living, organism. Everyone says
} } that s/he already knows this!
} }
} } Metachromasia is part physical chemistry, part biochemistry (and
} by
} } reason of those two, for many, little more than pure magic!).
} Metachromasia
} } has been extensively used, extensively studied and extensively reported
} } upon. Several items of information are clear [This may be an
} } exaggeration!].
} }
} } 1. The polychrome effect noted in the use of several related
} dyes
} } of the thiazine (quinone-imine/methylene blue) family is based on the
} } polymerization of dye molecules and the distances between adjacent acidic
} } groups on the substrate.
} }
} } 2. A loss of the polychrome effect could be attributed to*:
} } a. decrease in the concentration of some
} tissue/substrate
} } component
} } b. decarboxylation [- or - esterification with
} alcohol?]
} } c. degradation of the substrate to diffusible
} constituents.
} }
} } [* see Pearse, A.G.E.(1985), Histochemistry, Theoretical
} and
} } Applied(4th Ed.), Vol II, Churchill Livingstone, NY,NY, pp701-710. ISBN:
} } 0-443-02997-0.
} }
} } 3. In order for the metachromatic color. for (Toluidine Blue O,
} } that is purple [red + blue]) to be observed, the appropriately spaced
} acid
} } moieties must be deprotonated, i.e. negative in charge. Otherwise, the
} } orthochromatic color is observed.
} }
} } In other words, in order for there to be metachromasia of a
} } particular component of the tissue, that component must have
} appropriately
} } spaced negatively charged groups.
} }
} } 4. At pH 7, most 'acid' moieties in biological systems are
} } deprotonated, and thus, negatively charged. The isoelectric points of
} most
} } macromolecules are in the vicinity of pH 5 [This IS an exaggeration!]
} Where
} } pH can determine dye binding, pH can be used to partition objects in the
} } specimen space along a pH gradient [also in Pearse, same pages, but see
} } Methylene Blue Extinction (MBE) methods in many compendia]. [MBE used to
} } distinguish among histologic 'acid' mucosubstances (GAG's, etc.).]
} }
} } 5. A buffer may stabilize substances, react with them, or
} promote
} } changes in them (i.e. oxidation).
} }
} } 6. A bottle of dye, used for 5 years to produce a result, may,
} at
} } some moment, STOP behaving as it should. Pearse mentions that one of the
} } reasons that metachromasia was such a confused subject prior to the
} 50-60's,
} } was due to the fact that so many studies failed to use pure dyes.
} }
} } RULE: If a dye fails to function as it should. Try a different batch or
} } make up a fresh batch, or purchase a new supply. I have found that a
} simple
} } 0.1% solution of the dye gives rise to regular, reproducible
} metachromasia
} } which survives dehydration in absolute ethanol, but never in 95% ethanol.
} } That having been said, when I have performed the Azure B, pH 4.0 for
} } ethanol-acetic acid(3:1)(Clark) fixed nucleic acids (Flax and Himes,
} 1952),
} } I follow their protocol for dehydration and use tertiary butanol.
} } Preparations I made in the mid-60's still show metachromasia, albeit with
} } some overall loss of color in my personally prepared Damar-xylene
} mountant.
} }
} } The questions about the failure of any regularly used dyeing
} } procedure amount to a scientific challenge that are best addressed by the
} } one who is having the problem. Since the variables are many, the sources
} of
} } failure are also many. One has to learn how to perform a component
} analysis
} } in order to efficiently address such a problem. Since much of what one
} does
} } in dyeing/staining is by prescribed protocol, it should be clear that if
} } anything has changed, it is the operator who is in the best position to
} DO
} } the troubleshooting. When I ran the Flax and Himes procedure, I always
} } retained the blocks of previously sectioned material. Each sectioned
} block
} } was dipped in paraffin to cover the exposed tissue and stored carefully
} } away. I was especially careful of those specimens, prepared for any
} } particular purpose, in the event I ever required a 'known' tissue source
} of
} } a good result. Even so, I was aware that the stored specimens would not
} be
} } the same in two years as those I sectioned yesterday. I learned this
} when I
} } addressed the issue of saving tissues labeled with tritiated thymidine.
} } Why, I asked parenthetically, should I be able to determine that loss of
} } radioactivity by disintegration was not going to be augmented by loss due
} to
} } progressive destruction of DNA, if I had no specific knowledge of how
} much
} } DNA/nucleus/section was present in the starting material? So, I learned
} } that I would not be able to have compete faith, even in the best of my
} } archived specimens, 5 or 10 years in the future.
} }
} } NOTE: Acridine orange was one of the first fluorescent dyes
} used by
} } virologists in the late 50's/early 60's. One was able to distinguish
} } between single stranded RNA and double-stranded DNA until someone noted
} the,
} } then recent, discovery of the RNA viruses. "Nuts!" Even now, there is
} an
} } extensive literature on the use of AO fluorescence for double-dyeing the
} } nucleic acids in cell nuclei.
} }
} } An absolute obligation of old windbags is a SUMMARY: If there
} is a
} } single point in all of this, it is this. In the application of
} } metachromasia there are considerations of mass action, pH, purity of the
} } dye, and treatments preceding and following the application of the dye.
} If
} } each adds an order of magnitude to the number of variables involved,
} there
} } are 4-5 such ordinal magnifications through the process. If the physical
} } and chemical bases of histologic methods are understood at less than
} optimal
} } levels, then troubleshooting will be a problem that has little hope for
} } success. On the other hand, even one who has only the recipe to which
} s/he
} } can refer can be taught component analysis of that recipe.
} }
} } Mom used to say, "If you don't know what kind of flour to put in
} } your bread, try any flour and the bread will let you know if you were
} right
} } or wrong. If you were wrong, and really want bread, then you will have
} to
} } try another flour, and another, until the bread tells you that you are
} } finally right. Just don't change any other part of the recipe while you
} are
} } testing the flour." Ah! If only more of us had learned to bake when we
} } were young.
} }
} } END: Respectfully submitted,
} }
} } Fred Monson
} }
} }
} } Frederick C. Monson, PhD
} } Center for Advanced Scientific Imaging
} } Schmucker II Science Center
} } West Chester University
} } South Church Street and Rosedale
} } West Chester, Pennsylvania, USA, 19383
} } Phone: 610-738-0437
} } FAX: 610-738-0437
} } fmonson-at-wcupa.edu
} } CASI URL: http://darwin.wcupa.edu/casi/
} } WCUPA URL: http://www.wcupa.edu/
} } Visitors URL: http://www.wcupa.edu/_visitors/
}
} _____________________________________
}
} Sergey Ryazantsev Ph. D.
} Electron Microscopy
} UCLA School of Medicine
} Department of Biological Chemistry
} Box 951737
} Los Angeles, CA 90095-1737
}
} Phone: (310) 825-1144
} FAX (departmental): (310) 206-5272
} mailto:sryazant-at-ucla.edu
}
}
}
}


From daemon Mon Jun 3 11:50:25 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Mon, 3 Jun 2002 12:43:30 -0400
Subject: Legal Image Standards???

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,

I don't know much about this other than it exists as an issue in government
circles such as those found revolving at FDA and NCI.

URL: http://www3.cancer.gov/scienceresources/announcements/imaging.html

URL: http://www.acuotech.com/fda_filing.asp

Interesting Commercial Site: http://www.archivebuilders.com/

and also see:
http://www.archivebuilders.com/whitepapers/22041p.pdf

What constitutes a 'legal' digital, archived image has become a VERY
interesting topic.

Have a good one,

Fred

Frederick C. Monson, PhD
Center for Advanced Scientific Imaging
Schmucker II Science Center
West Chester University
South Church Street and Rosedale
West Chester, Pennsylvania, USA, 19383
Phone: 610-738-0437
FAX: 610-738-0437
fmonson-at-wcupa.edu
CASI URL: http://darwin.wcupa.edu/casi/
WCUPA URL: http://www.wcupa.edu/
Visitors URL: http://www.wcupa.edu/_visitors/


From daemon Mon Jun 3 12:23:13 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Mon, 3 Jun 2002 13:12:15 -0400
Subject: Jeol 100C

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We have a JEOL 100C top entry scope for sale to anyone interested. Has 2
sets of plate film boxes and holders. Asking $10,000 or Best offer with
removal and shipping at buyers labor and cost. Scope has been under service
contract for the last 10 years.

Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University
609-258-5432

Visit us at http://www.molbio.princeton.edu/facility/confocal/index.html



From daemon Mon Jun 3 12:29:42 2002



From: Mike McKay :      mike.mckay-at-vitana.com
Date: Mon, 3 Jun 2002 13:21:06 -0400
Subject: RE: TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ms. Clarkson,

I represent Vitana Corp., a digital camera manufacturer based in Ottawa,
Canada. We manufacture the PixeLINK brand of megapixel FireWire cameras
aimed at the scientific and industrial imaging markets.

At $1695 USD, our monochrome 1.3 megapixel camera may be of interest to you.
PixeLINK cameras provide excellent value and could act as a megapixel
preview system in support of a higher-end camera. We also manufacture a
color version priced at $1795. Please visit our web site
(http://www.pixelink.com) for more detailed information.

To answer your questions,

* What make & model camera system do you have?

The PixeLINK PL-A641 Monochrome camera is a 1.3 megapixel camera, connected
to a computer with a single FireWire cable. There are no other cables,
framegrabbers, nor power supplies. The camera uses a CMOS sensor and is
well suited to brightfield applications. The camera body mounts to the
photomultiplier or video coupler via a standard "C"-mount. For more
information on the cameras, please visit the product page on our web site -
http://www.pixelink.com/products/600.htm.

* Do you feel you get micrograph-quality resolution and images
utilizing that system?

Uncooled and with 1.3 megapixels and 10 bit sensitivity, the PixeLINK
cameras will not provide comparable images to photographic systems. The
comparison would have to be in terms of cost, speed and ease of use.

* Ease of use--is it user-friendly or cumbersome?

PixeLINK products are designed to be user-friendly. The camera equipment
and software can be installed and functional in just a few minutes. The
PixeLINK Capture application allows full control of the image capture
process. Exposure, brightness, and gamma are all easily adjusted.
Darkfield and brightfield corrections can also be applied. Images can be
saved, sequential numbered, in TIFF or bitmap formats. A bitmap overlay can
be applied to the image. Time-lapse operation is possible and the camera
can be operated with a remote control foot switch.

The most useful benefit to the user is the real-time preview. A full-screen
preview (1280x1024) at up to 14 frames per second can be used to observe the
specimen prior to image capture. At the more common VGA resolution
(640x480), the speed increases to 30 frames a second. At these speeds,
there is little need to examine the image through the eyepiece of the
microscope. The result in increased productivity for the operator.

* Ease and/or quality of service--are any of the components
serviceable by you? How quickly & easily is it to get a service rep?

The camera requires little support or service. Installation is simple and
operations training is not required. For support, call us at 1-800-4VITANA
(1-800-484-8262).

I'd be happy to answer any other questions you may have.

Yours,

Michael McKay

{ { { { { { { { { { { { { { { { { { { { { { { P i x e L I N K } } } } } } } } } } } } } } } } } } } } } } } } }
Michael McKay, Product Manager {mailto:mike.mckay-at-vitana.com}
Vitana Corporation
2500 Don Reid Drive Tel: (613) 247-1211 x 152
Ottawa, Ontario Cell: (613) 859-6174
Canada K1H 1E1 Fax: (613) 247-2001
"Making Digital Imaging Simple {http://www.pixelink.com} "
{ { { { { { { { { { { { { { { { { { { { { { { P i x e L I N K } } } } } } } } } } } } } } } } } } } } } } } } }



From daemon Mon Jun 3 12:33:16 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Mon, 3 Jun 2002 13:23:37 -0400
Subject: Jeol 100C

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We have a JEOL 100C top entry scope that is looking for a new home for
anyone interested. It has 2 sets of plate film boxes and holders. Asking
$10,000 or BO with removal and shipping at labor and cost. Scope has been
under service contract for the last 10 years.

Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University
609-258-5432

Visit us at http://www.molbio.princeton.edu/facility/confocal/index.html



From daemon Mon Jun 3 12:50:23 2002



From: Mary Gail Engle :      mgengle-at-pop.uky.edu
Date: Mon, 03 Jun 2002 13:40:28 -0400
Subject: Re: Ultracut Fluorescent bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have a Reichert Ultracut E and it takes a Philips TL 4W/33 XD6 or an
OsramL 4W/ 25, Weiss-universal-white. You might try Bulbman,
1-800-648-1163, or www.bulbman.com if you can't find them locally.
Mary Gail Engle

At 09:42 AM 6/3/02 -0500, Greg Strout wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Mary Gail Engle
Sr. Research Laboratory Manager
Electron Microscopy & Imaging Facility
Health Sciences Research Bldg. 001
University of Kentucky
Lexington, KY 40536-0305

phone 859-323-6108
fax 859-257-9700


From daemon Mon Jun 3 12:53:40 2002



From: Jeff Thompson :      jthompso-at-csusb.edu
Date: Mon, 03 Jun 2002 10:45:13 -0700
Subject: Auto Film Processor for EM negatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are evaluating an automatic film processor for developing EM
negatives from AFP Imaging Corp. (Mini-Medical series). We are trying
to develop SO-163 EM negative film with D-19 developer. No problem when
we develop manually in a tank. In the new processor, the film comes out
foggy/milky in a botchy pattern at most temperatures with a development
time of 130 seconds. If the temperature is maintained at exactly 85.6 F
then the film is clear. A second problem is that we see what we think
are roller marks on the film. A final note: the processor works fine
with X-ray film (Kodak-AR) for use in autoradiography.

My questions: Is anyone using a automatic processor for EM negatives
successfully?
Do you have any suggestions for tweaking the AFP processor?
Is there something special about the emulsion and/or gelatin coating on
the SO-163 film?

Thanks.

Jeff Thompson
Director, Electron Microscope and Image Analysis Center
Department of Biology
California State University
San Bernardino, CA 92407
909-880-5315



From daemon Mon Jun 3 13:20:32 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Mon, 3 Jun 2002 14:13:35 -0400
Subject: FDA and Electronic Documents

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Afternoon Listers,
I thought that the information/sites listed herein might be of
interest to many of you, so here it is.

I have been aware of the issue of "acceptability" in electronic
image/document storage and archiving. What I have not done until now is
search out the sources of useful information so that the issue comes into
greater focus, so to speak. While the sites listed below are United States
Government sites, the issues, as seen by the government regulators, are
exposed, at least in part. The most important thing is to note the
complexity of the issues as summarized in the first site. Hope this is a
welcome sort of sharing.

See a long summary of the FDA guidelines/requirements at:

URL: http://www.devicelink.com/mddi/archive/99/05/009.html

FDA Rule: 21 CFR 11 published on 20 Mar, 1997 is the relevant publication.
It considers more than just images.

At the FDA site of the Center for Devices and Radiologic Health, one can
read: "Guidance for the Submission Of Premarket Notifications for Medical
Image Management Devices"

URL: http://www.fda.gov/cdrh/ode/guidance/416.html (under "Diagnostic
Imaging" or "Picture Archiving and Communications Systems (PACS)") in the
following FDA Index page.
URL:
http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfTopic/topicindex/topindx
.cfm
or for the PDF: http://www.fda.gov/cdrh/ode/guidance/416.pdf
or for the printed document: call: 1-800-899-0381 and ask for
document 416.


I will know if this is OK with most listers if I don't receive a torrent of
negative responses.

Regards,

Fred Monson

Frederick C. Monson, PhD
Center for Advanced Scientific Imaging
Schmucker II Science Center
West Chester University
South Church Street and Rosedale
West Chester, Pennsylvania, USA, 19383
Phone: 610-738-0437
FAX: 610-738-0437
fmonson-at-wcupa.edu
CASI URL: http://darwin.wcupa.edu/casi/
WCUPA URL: http://www.wcupa.edu/
Visitors URL: http://www.wcupa.edu/_visitors/


From daemon Mon Jun 3 13:38:56 2002



From: Alan Nicholls :      nicholls-at-uic.edu
Date: Mon, 03 Jun 2002 13:34:37 -0500
Subject: Re:TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bill

What about the cost of film and chemicals over the lifetime of the
microscope? As a one off charge your dept chair was happy but the
additional costs of dealing with film are not small. Also as a totally
digital materials science lab the main additional benefit is being able to
see immediately the image and know it is in focus etc without wasting film
and time.

Alan

At 09:08 AM 6/3/2002 -0400, William Oxberry wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Alan W Nicholls, PhD
Director of Research Service Facility (Electron Microscopy)
Research Resources Center - East (M/C 337)
Room 100 Science and Engineering South Building
The University of Illinois at Chicago
845 West Taylor St
Chicago, IL 60607-7058

Tel: 312 996 1227
Fax: 312 996 8091
Office: Room 110

Web site www.rrc.uic.edu



From daemon Mon Jun 3 14:04:39 2002



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Mon, 3 Jun 2002 19:57:03 +0100 (GMT Daylight Time)
Subject: Re: Ultracut Fluorescent bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On my Ultracut E the writing is:
OSRAM (manufactuerer) L 4W/25

Interestingly my spare (unneeded in 13yrs here) is:
OSRAM L 4W/23

Dave


On Mon, 03 Jun 2002 09:42:41 -0500 Greg Strout
{gstrout-at-ou.edu} wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Does anyone know what the specs are for replacement fluorescent bulbs on
} a Reichert Ultracut microtome? The specs are not in the manual and our
} bulbs are so old that the printing is illegible. TIA.
}
} --
} ==================================================================
} Greg Strout
} Electron Microscopist, University of Oklahoma
} WWW Virtual Library for Microscopy:
} http://www.ou.edu/research/electron/www-vl/
} e-mail: gstrout-at-ou.edu
} Opinions expressed herein are mine and not necessarily those of
} the University of Oklahoma
} ==================================================================
}
}
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



From daemon Mon Jun 3 14:33:24 2002



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Mon, 3 Jun 2002 20:26:27 +0100 (GMT Daylight Time)
Subject: Re: LM Need cutting wheel specs for LKB Histo Knifemaker 2078

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In the UK we can get LKB spares from Leica. Perhaps they
can help.

Dave


On Mon, 3 Jun 2002 09:52:29 -0400 (EDT)
"sjb27-at-cornell.edu"-at-sparc5.microscopy.com wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Recently received a used LKB Histo Knifemaker 2078
} (makes Ralph glass knives, not triangular glass knives)
} but need to replace the cutting/scoring wheel. The
} manual does not list the specifications of the wheel,
} and the company that made the Knifemaker is no longer
} in business making knifemakers.
}
} Does anyone know the specs, or know a supplier for the
} cutting wheels?
}
} Thanks,
} Sandra Borgardt
} L. H. Bailey Hortorium
} 462 Mann Library
} Cornell University
} Ithaca, NY 14853
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



From daemon Mon Jun 3 19:48:18 2002



From: Mark YEADON :      m-yeadon-at-imre.org.sg
Date: Tue, 4 Jun 2002 08:43:43 +0800
Subject: Postdoc, In-situ TEM, Growth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Colleagues,

I would be grateful if you could bring this job ad to the attention of
anyone whom you think may be interested.

Many thanks,

Mark

Research Associate: In-situ TEM, Crystal Growth

The Institute of Materials Research and Engineering (IMRE) is seeking a
Postdoctoral Research Associate with a strong background in transmission
electron microscopy. The candidate will be involved in the in-situ growth
and characterisation of thin films and nanostructured materials using the
newly commissioned MERLION system, a modified JEOL 200kV TEM with ultrahigh
vacuum column. The system is equipped with in-situ electron beam
evaporators and gas injectors enabling real-time in-situ observations of
materials growth. The system is also equipped with a Gatan Image Filter
(GIF) and Dualview camera system.

IMRE is supported by the Singapore Agency for Science, Technology and
Research (A*STAR) and is situated on the campus of the National University
of Singapore (NUS). (www.imre.org.sg; www.nus.edu.sg).

Interested candidates should submit a comprehensive CV/resume together with
the names of at least 3 referees, to the address below. Email
correspondence is encouraged, and further details can be supplied upon
request.

=================
Dr Mark Yeadon
Institute of Materials Research and Engineering
3 Research Link, Singapore 117602

Tel: (+65) 6874 8591
FAX: (+65) 6872 0785
Email: m-yeadon-at-imre.org.sg
=================

%%%%%%%%%%%%%%%%%%
Mark Yeadon

President, Microscopy Society of Singapore

Senior Research Fellow
Institute of Materials Research and Engineering
3 Research Link
Singapore 117602

Assistant Professor
Department of Materials Science
National University of Singapore
Singapore 119260
http://www.matsci.nus.edu.sg/STAFF/Mark.html

TEL: (+65) 6874 8591
FAX: (+65) 6872 0785
Email: m-yeadon-at-imre.org.sg



From daemon Mon Jun 3 23:10:52 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Mon, 03 Jun 2002 21:02:18 -0700
Subject: Re:TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Using digital camera (600W BioScan, Gatan) with TEM over last couple of
month I find the following.

-most people prefer to use digital imaging even we offer equal access to
the film as well. It's because in most cases people need to evaluate
sample quality etc, they simply don't need film quality. From another hand
they could use film anytime if they wanted.

-with digital images you have chance to see your sample immediately
(important for evaluation/students in rush/deadlines etc).

-If you do not enlarge the digital image, 7x7" draft printout on laser
printer (standard) is looks very similar to the image produced from scanned
film. Manipulating with image resolution, dpi etc, you should keep in
mind, that 1200 dpi printer resolution mean that printer could produce 1200
dots per inch in the row making solid line. In order to create shadows of
gray, printer should print dots with space. For instance, 50% of grey
would be represented by 'dot-space' sequence. It mean, that resolution
would be 50% from 1200 = 600 dpi. For 30% gray, it would be 400 dpi. So,
there is no reason to abuse printer sending 1200 dpi 50 Mb image. On
practice, I could not recognize difference between 1x1K at 144 dpi digital
camera image from scanned from film 1200 dpi image: On laserJet (similarly
on our Tektronix dye-sub) - they are looks very the same. BUT: the
difference is that you could enlarge 1200 dpi image (and print with the
same quality) and you COULDN'T do so with 1x1K image from digital
camera. Another remark here: in 'prestigious' magazines like
Science/Nature technical editors have tendency to reduce image to the
postal stamp size, so you could do it by yourself and there is no need to
use film for such small images, digital camera will work just fine!

-another advantage of the digital camera, which, actually, I did not
expect, is its sensitivity. My camera is at least x10 more sensitive than
film, so exposure time is 0.1-0.2 sec versus my usual 1.5-2 sec for the
film, So, less drift and sample damage.

-another things are educational. Instead keep students in the dark and
teach them how to adjust binoculars (instead doing actual EM), we are
comfortably sitting around the 19" flat screen monitor with lights ON (and
nice, not loud classic music). I find it's more convenient to show how to
focus image on the computer screen rather than make a 'focus series', then
develop/print it. At this point students usually don't remember how image
was looks like on the microscope's screen and we have to start again. I
don't mean that students don't need to know how to operate
microscope/binoculars, I just mean that digital camera makes this process
more enjoyable and creative.

-another thing is academic: The live image from digital camera I broadcast
to the Departmental network, so you could see live image on any
departmental computer. PIs now comfortably sit in their offices and enjoy
watching how their students working on the TEM. Similarly it works for
Internet. We did a few session when our collaborators observe their
samples sitting far-far away from UCLA.

So, the bottom line is that TEM digital camera is very convenient
'supplement' to your EM. It does not replace the film, but enhance your
microscope's abilities. It's like power-steering in the car: convenient,
but not necessary - you may park your car in NY without this 'supplement'
right? As a matter of fact, cars without power-steering is cheaper.

Sergey

P.S. I just figured out that this message may be recognized as an
'advertisement'. So, it looks like I was working hard writing on ESL for
manufacturers. I would like to declare openly - I did not intend to do so
and nobody from Gatan or other manufacturer offered to me dinner or coffee
when I was writing it. I did it in sincere believe of what I was writing.

At 11:34 AM 6/3/02, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Tue Jun 4 03:54:27 2002



From: Janos Labar :      labar-at-mfa.kfki.hu
Date: Tue, 4 Jun 2002 10:45:29 +0200
Subject: Evaluation of SAED ring patterns from nano-phase materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear List Members,

I am happy to inform you that a brand new version (V2.0.7) of my free
program is now available at the same web-site where the previous version was
published
(www.mfa.kfki.hu/~labar/ProcDif.htm ).

Beside the new Multiple Document User Interface, new functionality has also
come: Some of the new features:

1/ You can generate Marker files on your own for known structures.
The new Structure Definition Module allows you to enter the
crystallographic data for known phases.
Automatic listing of the possible choices and some checking of the
consistency of the entered data are provided by the program.
(e.g. only the Space Groups in accordance with the selected Bravais
lattice are listed, Wyckoff positions for the selected Space Group are only
listed, and
a warning is given if the specified atomic coordinates are conflicting
with the multiplicity of the selected Wyckoff symbol)

Kinematic electron scattering (powder diffraction) data are calculated
to be used as Markers for the measured ring-patterns
(in contrast to using only X-ray data from a database, which is still a
valid option if the structure of the pahse is not known).

2/ You can copy the curves of the processed SAED patterns into 5 different
"Compare Memories". These data can be saved and later re-loaded.
Using this feature, you can compare distributions of intensities at any
time, without a need for later re-processing of previously processed SAED
patterns.

Compared distributions can be normalized on display to the same
full-scale.

3/ XRD data can also be loaded into Compare Memories for comparison with
SAED data. For this, the XRD data should be saved in a text file with the
following format:
2 columns, separated by coma. The first column is 2-Theta and the second
is Intensity (Counts).
A special 3-line header must be added:
"[Source = XRD]"
"[X-axis type = 2Theta]"
"[Lambda = 1.541]"
Obviously, the value of the wavelength should correspond to the true
value used in the XRD.
The header-format must be strict (case-sensitive,
character-by-character).

4/ New command buttons help stretch, compress and shift distributions in
either X or Y directions. The more accurate setting-possibility of the axes
still persists.

5/ New image formats are supported:
TIF (uncompressed) and RAW.
Both of them can be either 8-bit or 16-bit. Obviously, the point is in
using the 16-bit format, since the 8-bit format does not offer more that the
BMP-files.

6/ A simple Document is added to facilitate writing reports. It can save
text and pictures in Rich Text Format (.RTF files)
that can be read e.g. by MS Word for more sophisticated formatting.
Be aware that RTF files with embedded pictures are huge and saving them
takes a very long time.

{Alt} {Prt SC} can be used to copy shots of screen onto the Clipboard
from where you can insert them into the Document by {Ctrl} V.

Also see the new menu points for saving or copying graphs.

7/ Processing of negatives is done in two steps. First Select menu Process /
Use Inverted ... (this is a toggle-switch), then the usual Process /
Calculate Distribution.

8/ Option menu point helps you to determine
- where to find and store different types of files
- whether to obtain hints at every stage of your work
- what line-thickness to use in Schematic half-circles
- which minimal intensity is to be used to display Markers (later
versions will contain a possibility to show forbidden lines, too).

Some Folders are created for you as a suggestion where to store SAED
patterns, Structures, XRD Markers, Electron Diffraction Markers (EDM), etc.
You can change this as you wish. The program warns you if a non-existent
Path is specified as Default and offers the usual Windows interface
to select a new one, instead.

9/ Letter size is increased to produce better material for printing or for
Projected Presentations.

10/ Peak Search is separated with the inclusion of the possibility to change
the filter-parameters for search.
The window comes up with suggested parameters, but you can change them
to experiment how to find most peaks with less noise.
Peak positions hopefully became more accurate, but do not underestimate
the importance of visual inspection.

11/ I hope the program became more robust. However, I would appreciate
receiving any comments, suggestions or reports on errors.

I wish you good luck in using this program.

Best regards:

Janos L. Labar




From daemon Tue Jun 4 06:14:39 2002



From: Susan Carbyn :      CarbynS-at-agr.gc.ca
Date: Tue, 04 Jun 2002 07:05:26 -0400
Subject: Re: Ultracut Fluorescent bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Greg,

Our Reichert Ultracut E Microtome has Philips TL4W/33 F7 bulbs in it. Sounds like your bulbs lasted a long time, since you can't read the print. We haven't had to replace ours yet, so I guess they must last awhile.

Susan

Susan Carbyn
Electron Microscopy Technician
Atlantic Food and Horticulture Research Centre
Agriculture and Agri-Food Canada
32 Main Street, Kentville, N.S. B4N 1J5
Canada

Phone (902) 679-5535
Fax (902) 679-2311

E-Mail: carbyns-at-em.agr.ca

} } } Greg Strout {gstrout-at-ou.edu} 06/03/02 11:42AM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Does anyone know what the specs are for replacement fluorescent bulbs on
a Reichert Ultracut microtome? The specs are not in the manual and our
bulbs are so old that the printing is illegible. TIA.

--
==================================================================
Greg Strout
Electron Microscopist, University of Oklahoma
WWW Virtual Library for Microscopy:
http://www.ou.edu/research/electron/www-vl/
e-mail: gstrout-at-ou.edu
Opinions expressed herein are mine and not necessarily those of
the University of Oklahoma
==================================================================






From daemon Tue Jun 4 06:22:51 2002



From: LLOYD, PAMELA F [AG/1000] :      pamela.f.lloyd-at-Monsanto.com
Date: Tue, 4 Jun 2002 07:16:40 -0500
Subject: Ultracut Fluorescent bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Heather
A simple holder to make would be a flat piece of 4-6mm aluminium
same size as the slide or a little larger, drilled M4 in the centre,
fitted with non-ferrous metal spring clips similar to LM stage clips
Chris

Date sent: Fri, 31 May 2002 15:55:04 -0500 (CDT)
} From: Heather A Owen {owenha-at-csd.uwm.edu}
To: MSA Listserver {Microscopy-at-sparc5.microscopy.com}


Greg,

We have a Reichert/Leica Ultracut S. The fluorescent bulb is a 5W Osram
Dulux S/E. We get it from Mager Scientific Inc. in Dexter, MI. The part
number is 870041.

I don't know if this is interchangeable with the Ultracut E.

Pam Lloyd


Pamela F. Lloyd
Research Associate
Monsanto Co.
800 N. Lindbergh Blvd.
U1E
St. Louis, MO 63167
Phone: (314)694-6527
FAX: (314)694-8065
e-mail: pamela.f.lloyd-at-monsanto.com

-----Original Message-----
} From: Greg Strout [mailto:gstrout-at-ou.edu]
Sent: Monday, June 03, 2002 9:43 AM
To: Microscopy listserve


Does anyone know what the specs are for replacement fluorescent bulbs on
a Reichert Ultracut microtome? The specs are not in the manual and our
bulbs are so old that the printing is illegible. TIA.

--
==================================================================
Greg Strout
Electron Microscopist, University of Oklahoma
WWW Virtual Library for Microscopy:
http://www.ou.edu/research/electron/www-vl/
e-mail: gstrout-at-ou.edu
Opinions expressed herein are mine and not necessarily those of
the University of Oklahoma
==================================================================




From daemon Tue Jun 4 08:01:12 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Tue, 4 Jun 2002 08:54:29 -0400
Subject: RE: Mitotic index - McAuliffe cont'd

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree with Geoff, Augustin. Br-UdR is the current way to go. It actually
requires more work than 3H-TdR, but for short-term experiments, its use
avoids a LOT of safety and regulatory hassles. Problem is that in longer
term experiments, where 3H-TdR is known, by literature support, to 'breed
true', there appears to be less such support for long-term Br-UdR. If you
just count the layers (i.e., the number of steps to finished product),
3H-TdR comes out way ahead in resolution, ease of use and precision, BUT the
hassle in use of radioactive labels is almost totally limiting if one only
considers the added cost of handling the remains.

When I used 3H-TdR on rabbit urinary bladder in situ(vivo) in the early
90's, I ended up using very low dosage per gram of body weight (~0.1uCi/g)
when I introduced the label i.v. and normal (for me!) dosage (0.5uCi/ml or
g) in aerated Hank's BSS (NO BSA added!!) for in vitro incubations of either
whole bladder or bladder strips. Labeling indices proved to be similar in
three categories: whole bladder labeled either in vivo(1) or in vitro(2)
and bladder strips in vitro.

NOTE: the throw-away radiation level was 0.05uCi/g body weight at that time
(early '90's) so even when dosed less and exposed in a special manner,
hazardous waste disposal was still required for the carcass. Needless to
say, the growing cost of disposal caused me to expend both energy and time
in developing a means by which I could dose with 0.049uCi/g body weight and
not have to wait for a generation (mine!) for the autoradiograms to expose.
Got close, but time and money ran out.

Regards,

Fred Monson

Frederick C. Monson, PhD
Center for Advanced Scientific Imaging
Schmucker II Science Center
West Chester University
South Church Street and Rosedale
West Chester, Pennsylvania, USA, 19383
Phone: 610-738-0437
FAX: 610-738-0437
fmonson-at-wcupa.edu
CASI URL: http://darwin.wcupa.edu/casi/
WCUPA URL: http://www.wcupa.edu/
Visitors URL: http://www.wcupa.edu/_visitors/

} ----------
} From: Geoff McAuliffe
} Sent: Monday, June 3, 2002 4:58 PM
} To: Agustín Venzano
} Cc: HistoNet Server
} Subject: Re: Mitotic index
}
} Augustin:
}
} I believe that bromodeoxyuridine is a more accurate index of mitotic
} activity, but I can't remember the citation right now. Actually seeing
} mitotic
} figures is difficult, they don't last long! S-phase often lasts 10-12
} hours,
} mitosis lasts 20 minutes so incorporation of something into the DNA during
} S-phase is the way to go. The "old" way with tritiated thymidine
} autoradiography
} has problems due to disposal of animals and contaminated reagents.
}
} Agustín Venzano wrote:
}
} } Dear netters: I'm planning a sampling of ruminal papillae (i.e. special
} } structures endowed with epithelium and lamina propria located in the
} largest
} } forestomach of ruminants) in young cattle. The project is aimed at
} defining
} } the growth rate of these structures specialised in nutrients absorption,
} so
} } it is necessary to estimate the mitotic index. My questions are:
} }
} } 1.What staining would you prefer for seeing DNA and mitosis?
} } 2. Do you consider c-kit to be an adequate marker of the mitotic rate
} } through IHC in paraffin blocks?
} }
} } Thank you in advance
} }
} } Sincerely yours
} }
} } Agustin Jose Venzano Halliburton
} } DVM-Pathology Group
} } INTA, Argentina
}
} Geoff
} --
} **********************************************
} Geoff McAuliffe, Ph.D.
} Neuroscience and Cell Biology
} Robert Wood Johnson Medical School
} 675 Hoes Lane, Piscataway, NJ 08854
} voice: (732)-235-4583; fax: -4029
} mcauliff-at-umdnj.edu
} **********************************************
}
}
}
}
}


From daemon Tue Jun 4 08:15:00 2002



From: Awbrey, Donald :      DonaldAwbrey-at-texashealth.org
Date: Tue, 4 Jun 2002 08:07:21 -0500
Subject: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Dear EM Netters,

Does anyoun out there have experience with imaging plate technology (i.e.
Ditabis)? I
would like to hear pros and cons concidering this technology. How does it
compare to
digital CCD camera systems (cost, time, supplies, reliability. We are
looking into possibly
retrofitting our JEOL 100CX with an off-line system.


Thanks in advance


Donald G. Awbrey, HT (ASCP), QIHC
Electron Microscopy / Image Analysis
817-878-5647
donaldawbrey-at-texashealth.org



From daemon Tue Jun 4 08:34:06 2002



From: Chuck Butterick :      cbutte-at-ameripol.com
Date: 5/31/02 2:33 PM
Subject: TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We upgraded an old Philips 300 with an AMT system with a Hamamatsu
C4742-95 camera 3.5 years ago. The system was mounted in the old 35mm
camera port, which does not have the resolution an under-the-column
camera will have, because I didn't want to lose the capability of
using film. The system has allowed a superior throughput of work with
same day turnarounds on projects that took 2 weeks before. Further,
digital image acquisition permits image analysis without any
intervening steps, which is crucial, for example, to calculating the
particle size of carbon black aggregates in a time-conscious
production environment.

For every day work, a laser printer works just fine. When printing
images for show/publication, an inkjet printer using a high quality
inkjet paper (Kodak or HP photo quality papers are my choice) does
very well. The ability to make high quality enlargements, though, is
restricted. Wet chemistry, film and paper is better, but not by
much...and, arguably, may not be worth the extra time and effort.

Not only does a digital system save on film, paper and chemistry (with
all the associated environmental concerns), it saves on
labor/time...and that is where the largest dollar savings is. Further,
as happened in my situation, I didn't need a darkroom any more so that
room was converted to house the new SEM/EDX system, which is a more
profitable use of the square footage. Even academic chairman (well,
in medical schools anyway) are concerned about the amount of research
dollars per square foot.

The choice of a system must always factor in service as well as
capability and price. AMT and Gatan both have excellent systems.
However, no one, given my circumstances at the time, would have chosen
an alternate system. Given another chance to buy a digital system for
a TEM, the AMT folks would have first shot, with the alternates having
an uphill, but not impossible, battle to convince me otherwise.

AMT competitors make excellent products. I'm just very satisfied with
cost, service, and instrumentation obtained from AMT. The usual
disclaimers apply.

Chuck Butterick
Degussa Corporation
Borger, TX


______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


We are looking to purchase a digital camera system for our TEM and
would like some feedback from those of you who already utilize such systems.
Some questions I'd like to ask are:

* What make & model camera system do you have?
* What type, make & model printer(s) do you use, (e.g. inkjet,
dye-sublimation, silver halide)?
* Do you feel you get micrograph-quality resolution and images
utilizing that system?
* Ease of use--is it user-friendly or cumbersome?
* Ease and/or quality of service--are any of the components
serviceable by you? How quickly & easily is it to get a service rep?
* How long have you had this system?
* Would you recommend this system--why or why not?
* What, if anything, might you have done differently?

I'd greatly appreciate any responses. Thank-you in advance for your
help.

Donna R. Clarkson





From daemon Tue Jun 4 08:58:11 2002



From: Anthony J. Garratt-Reed :      tonygr-at-mit.edu
Date: Tue, 04 Jun 2002 09:51:44 -0400
Subject: Re:TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id IAA06937
for dist-Microscopy; Tue, 4 Jun 2002 08:57:05 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id IAA06930
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Tue, 4 Jun 2002 08:56:34 -0500 (CDT)
Received: from pacific-carrier-annex.mit.edu (PACIFIC-CARRIER-ANNEX.MIT.EDU [18.7.21.83])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id IAA06922
for {Microscopy-at-sparc5.microscopy.com} ; Tue, 4 Jun 2002 08:56:22 -0500 (CDT)
Received: from central-city-carrier-station.mit.edu (CENTRAL-CITY-CARRIER-STATION.MIT.EDU [18.7.7.72])
by pacific-carrier-annex.mit.edu (8.9.2/8.9.2) with ESMTP id JAA07301
for {Microscopy-at-sparc5.microscopy.com} ; Tue, 4 Jun 2002 09:51:45 -0400 (EDT)
Received: from melbourne-city-street.mit.edu (MELBOURNE-CITY-STREET.MIT.EDU [18.7.21.86])
by central-city-carrier-station.mit.edu (8.9.2/8.9.2) with ESMTP id JAA18909
for {Microscopy-at-sparc5.microscopy.com} ; Tue, 4 Jun 2002 09:51:45 -0400 (EDT)
Received: from emlab.mit.edu (EMLAB.MIT.EDU [18.82.0.121])
by melbourne-city-street.mit.edu (8.9.2/8.9.2) with ESMTP id JAA01398
for {Microscopy-at-sparc5.microscopy.com} ; Tue, 4 Jun 2002 09:51:45 -0400 (EDT)
Message-Id: {5.0.2.1.2.20020604093713.01985700-at-hesiod}
X-Sender: tonygr-at-hesiod (Unverified)
X-Mailer: QUALCOMM Windows Eudora Version 5.0.2


Alan is correct about the cost of film, etc. We charge our users 75˘US per
sheet for film to cover this cost. I disagree, though, about seeing the
image is in focus - if you have a digital system then you can use it to
check the image is in focus if you like, (and it certainly helps at very
high magnifications), but you can still record, with advantage, on film,
when the image will still be in focus.

The issue, though, goes beyond cost. For some users, digital imaging will
give excellent results, and meet all their needs. For others, the extreme
(by current standards) information density of conventional film is still
essential to get the job done. One has to be careful before deciding to
abandon an existing darkroom facility. Of course, the equation could be
different if a new facility is being designed. We still teach our users to
take all their images on film, for subsequent scanning. We do have digital
capture systems, but they are generally used only by outside users who
haven't learned how to handle film.

In changing from film to digital imaging, it is also important to
understand that significant re-education of users is important. The two
media have different characteristics, which must be understood and
accounted for if best results are to be obtained.

Tony.


At 01:34 PM 6/3/2002 -0500, you wrote:
} Bill
}
} What about the cost of film and chemicals over the lifetime of the
} microscope? As a one off charge your dept chair was happy but the
} additional costs of dealing with film are not small. Also as a totally
} digital materials science lab the main additional benefit is being able to
} see immediately the image and know it is in focus etc without wasting film
} and time.
}
} Alan


** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6478
**




From daemon Tue Jun 4 09:19:51 2002



From: Mike Marko :      marko-at-wadsworth.org
Date: Tue, 4 Jun 2002 10:04:48 -0400
Subject: TEM tomography workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Colleagues,

A workshop on electron tomography will be held on 14-16 November 2002
in Albany, New York. The workshop is being offered at the Wadsworth
Center laboratories by the Resource for Visualization of Biological
Complexity (RVBC), a Biotechnology Resource supported by NCRR/NIH.

The workshop is intended to provide hands-on experience in electron
tomography, with an emphasis on data collection. Several different data
collection procedures, software packages, and electron microscopes will
be used. There will be daily practical sessions. There will also be
lectures, demonstrations, and practical sessions covering alignment and
reconstruction, visualization techniques, and preparation of
frozen-hydrated specimens.

Instructors and lecturers will include RVBC staff, as well the
individuals directly involved in software and technique development for
the systems that will be used during the workshop.

There is no registration fee.

Registration deadline: 1 October, 2002.

For more information and to register, please go to:

http://home.nycap.rr.com/cdmms/workshop

Michael Marko
Workshop Coordinator
Wadsworth Center, Albany, NY
marko-at-wadsworth.org


From daemon Tue Jun 4 09:29:48 2002



From: William H Roberts :      William.H.Roberts-at-usa.dupont.com
Date: Tue, 4 Jun 2002 10:28:38 -0400
Subject: Re: Legal requirements for microscope images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings Listers,

Normally I am a lurker, a parasite if you will, gleaning valuable
information and insights from the listings without contributing. However,
the current thread concerning the legality of "altered" or "enhanced"
digital images stirs old feelings from my youth in the days when I was a
forensic scientist (we called ourselves criminalists in those days). The
issue that was hot then between the attorneys, both prosecution and
defense, and myself was over the question of my resistance to introducing
any images at all to bolster my testimony as an expert witness. I think
that some of my arguments are still cogent and might be appropriate to
reconsider now.

First, if a putative expert witness survives voir dire and is declared an
expert by the court (and this must be done every time expert testimony is
to be presented, although it does get easier as one's reputation is
established), then the witness is permitted to render opinions which is
something an ordinary witness is not allowed to do. These opinions are
allowed based on the expectation that the expert possesses knowledge or
skills beyond those of the average person in the subject area(s) wherein
his or her testimony will be rendered. The expert's testimony is not and
should not be expected to inflame, excite, or even to educate the court or
the jurors to the point where they would be able to draw their own
conclusions from the same data set that the expert worked from. Hence, the
expert should not supply spectra, images, graphs, etc. to support his or
her testimony in open court.

Second, as has been alluded to in an earlier posting, questions concerning
the validity of testimony in the form of challenges from opposing expert
witnesses may well be expected and serve the valuable cause of keeping us
all honest. I never released raw or enhanced images or data to defense
experts, only samples of the specimens I had examined. It does not forward
the cause of justice to have one expert influence the opinion of another by
sharing the results of analyses or images. There is also no reason to
inflate the profits of independent consultants by supplying them with work
results done at the taxpayers expense.

Third, an image is only a representation of reality. It is not that
reality in and of itself. Therefore, even the "raw" image cannot be deemed
to convey any absolute information or truth about a sample or a scene in
the sense that it is entirely free of any alteration, subjectivity,
distortion, or misleading appearances. Further, if an individual is so
unscrupulous as to present false testimony, whether with supporting imagery
or not, then any supposed record of image enhancement would be highly
suspect as well.

} From the foregoing I would argue that there is no basis for a set of
"legal" restrictions on what can or cannot be done to an image. An image
is just an abstraction of reality. In many cases the image does allow us
to observe or interpret reality in a way that we as humans would be unable
to do given the limitations of our innate sensory abilities. However, we
will always have to deal with the further reality that the skill levels and
conscientiousness of all individuals who generate images are not equal,
just as not every referee will call the same pitch the same way, for
whatever reason. We are humans and rules don't make us better or worse
photographers.

Just some ramblings submitted for your consideration.



From daemon Tue Jun 4 09:39:47 2002



From: Clarkson Donna R Contr USAMRD :      donna.clarkson-at-brooks.af.mil
Date: 5/31/02 2:33 PM
Subject: TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chuck,
Thank-you for your input. As with many others the opinions are mixed,
but seem to lean toward the AMT camera system. I am a little confused on one
issue, though. You mentioned that you have lost some resolution by
side-mounting the camera; others have noted that they thought they had lost
resolution by under-the-column mounting. Hmmmm, I guess it depends on the
scope it's mounted to.

Best regards,

Donna

Northrop Grumman Information Technology
for U S Army Medical Research Detachment
at Brooks Air Force Base
Phone (210) 536-1416
FAX (210) 536-1449
e-mail donna.clarkson-at-brooks.af.mil


-----Original Message-----
} From: Chuck Butterick [mailto:cbutte-at-ameripol.com]
Sent: Tuesday, June 04, 2002 8:15 AM
To: Microscopy-at-sparc5.microscopy.com; donna.clarkson-at-brooks.af.mil


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


We are looking to purchase a digital camera system for our TEM and
would like some feedback from those of you who already utilize such systems.

Some questions I'd like to ask are:

* What make & model camera system do you have?
* What type, make & model printer(s) do you use, (e.g. inkjet,
dye-sublimation, silver halide)?
* Do you feel you get micrograph-quality resolution and images
utilizing that system?
* Ease of use--is it user-friendly or cumbersome?
* Ease and/or quality of service--are any of the components
serviceable by you? How quickly & easily is it to get a service rep?
* How long have you had this system?
* Would you recommend this system--why or why not?
* What, if anything, might you have done differently?

I'd greatly appreciate any responses. Thank-you in advance for your
help.

Donna R. Clarkson




From daemon Tue Jun 4 10:02:12 2002



From: Clarkson Donna R Contr USAMRD :      donna.clarkson-at-brooks.af.mil
Date: Tue, 4 Jun 2002 09:55:53 -0500
Subject: Ultracut Fluorescent bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id KAA07314
for dist-Microscopy; Tue, 4 Jun 2002 10:01:14 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id KAA07307
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Tue, 4 Jun 2002 10:00:43 -0500 (CDT)
Received: from diamondback.brooks.af.mil (diamondback.brooks.af.mil [140.140.58.5])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id KAA07300
for {Microscopy-at-sparc5.microscopy.com} ; Tue, 4 Jun 2002 10:00:31 -0500 (CDT)
Received: from diamondback.brooks.af.mil (root-at-localhost)
by diamondback.brooks.af.mil with ESMTP id g54EtsE22702
for {Microscopy-at-sparc5.microscopy.com} ; Tue, 4 Jun 2002 09:55:54 -0500 (CDT)
Received: from ballys.brooks.af.mil (ballys.brooks.af.mil [140.140.249.203])
by diamondback.brooks.af.mil with SMTP id g54Etsg22698
for {Microscopy-at-sparc5.microscopy.com} ; Tue, 4 Jun 2002 09:55:54 -0500 (CDT)
Received: from sahara.brooks.af.mil ([140.140.249.207])
by ballys.brooks.af.mil (NAVGW 2.5.2.11) with SMTP id M2002060409542801502
; Tue, 04 Jun 2002 09:54:28 -0500
Received: by sahara.brooks.af.mil with Internet Mail Service (5.5.2653.19)
id {LFVKH3FZ} ; Tue, 4 Jun 2002 09:55:53 -0500
Message-ID: {7BE3CD36C319D311B5FF00902746460705F54F9B-at-harrahs.brooks.af.mil}


Greg,
We have the Ultracut R and S, in addition to the E. If you need bulbs
for either of these you can order directly from Osram Sylvania at
www.osram.com. The R and S models use a U-shaped bulb, Sylvania Dulux S/E,
5W compact fluorescent, # 20315. I hope this helps.

Sincerely,
Donna R. Clarkson

Northrop Grumman Information Technology
for U S Army Medical Research Detachment
at Brooks Air Force Base
Phone (210) 536-1416
FAX (210) 536-1449
e-mail donna.clarkson-at-brooks.af.mil


-----Original Message-----
} From: Greg Strout [mailto:gstrout-at-ou.edu]
Sent: Monday, June 03, 2002 9:43 AM
To: Microscopy listserve


Does anyone know what the specs are for replacement fluorescent bulbs on
a Reichert Ultracut microtome? The specs are not in the manual and our
bulbs are so old that the printing is illegible. TIA.

--
==================================================================
Greg Strout
Electron Microscopist, University of Oklahoma
WWW Virtual Library for Microscopy:
http://www.ou.edu/research/electron/www-vl/
e-mail: gstrout-at-ou.edu
Opinions expressed herein are mine and not necessarily those of
the University of Oklahoma
==================================================================




From daemon Tue Jun 4 10:20:24 2002



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Tue, 04 Jun 2002 10:13:39 -0500
Subject: Ge detector problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a Ge detector on our EDS system that has been there for several years.

Our normal procedure is to fill the dewar with LN2 twice a week (Monday or
Tuesday and then Thursday or Friday). We would like to think the dewar had
enough capacity to last a full week, and it used to. However, in the
end-of-the-week rush, I forgot to fill the dewar and came in to a warm
detector on Monday. It had been filled last Tuesday, so something seems to
be degrading that we no longer can go even 6 days on a fill.

We have an LN2 alarm, and it had worked and presumably cut off the high
voltage. We refilled the dewar, gave it an hour or more to cool off, then
tried the system again. We got a spectrum, but with artifacts.

I tried conditioning the detector (three times over 24 hours) and
recalibrating the electronics. The strobe peak got back into balance, but
artifacts still remain. I have posted three documents on the web at
ftp://www.marl.iastate.edu/Ge_detector/ for those that might be interested
in examining the problem for themselves. One file is a bitmap of a line
trace, another is a color bitmap, and the third, most detailed view, is a
figure embedded in a word document.

All three files show the same thing - an overlay of spectra from an
aluminum sample holder and a titanium standard. Both show the
characteristic peak but with perhaps more of a low energy tail. But the
real problem is a large hump in the background down-scale from the peak.
The hump follows the characteristic peak up and down the energy scale with
the element and always seems to peak at about 38% of the characteristic
energy.

I have been given one diagnosis and prognosis from the manufacturer. But I
wonder if the resident experts on this list might have a second opinion. I
welcome your comments, particularly if they might lead to a quicker,
cheaper fix than sending the detector back.

Warren

-------------------------------------------
No files should be attached to this message
-------------------------------------------
Warren E. Straszheim, Ph.D.
Materials Analysis and Research Lab
Iowa State University
23 Town Engineering
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
Web: www.marl.iastate.edu

Scanning electron microscopy, x-ray analysis, and image analysis of materials
Computer applications and networking




From daemon Tue Jun 4 10:52:32 2002



From: Menon Sarath K Contr AFRL/MLLM :      Sarath.Menon-at-wpafb.af.mil
Date: Tue, 4 Jun 2002 11:42:14 -0400
Subject: Microprobe with FEG

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Could experts comment on advantages and disadvantages of using a microprobe with a FEG vs. a LaB6 emitter?
Pointers to any recent publications will be nice too.

thanks

Sarath Menon
Scientist
UES Inc.
Dayton, OH


From daemon Tue Jun 4 12:06:09 2002



From: Eric Anderson :      anderson_e-at-southernct.edu
Date: Tue, 04 Jun 2002 12:53:46 -0400
Subject: Re: Philips EM400 Matching Mains Transformer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Greetings All!
}
} We are in the process of setting up a second-hand EM400 acquired
} recently, and have found that the mains matching transformer was not
} included. Our raw supply main is 237 volts/60Hz (118V on one leg,
} 119V on the other, stability unknown), and I'm thinking this is not
} close enough to the specified 220V to go without the transformer. Any
} ideas? If we do need some line conditioning, can anyone recommend a
} particular device, or source for the original Philips transformer?
}
} Many thanks for any tips!
} -Eric
} --
}
Eric Anderson
SCSU Physics Adjunct
203-392-6455
anderson_e-at-southernct.edu

}



From daemon Tue Jun 4 13:10:47 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Tue, 4 Jun 2002 12:10:10 -0600
Subject: Re:TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tony,

you are of course right, that the "extreme" information density of film
cannot be approached today with digital cameras. This is where the different
types of digital TEM cameras come into play. On most microscopes there are
two possibilities to attach a camera: the 35 mm port (side-mount) and below
the film chamber (bottom-mount). BOTH types, by the way, allow you to keep
the film camera in most cases.

The side-mount cameras usually "see" an area that is approximately the same
as the area on a negative. Of course the resolution of the digital image is
then worse than the resolution on the film. The bottom-mounted cameras
usually "see" a much smaller area (plus there may be an addition
magnification due to geometric reasons), but they normally have a resolution
that is similar to film.

In our experience (and we have been producing these cameras for many years
now), there is almost never anybody who really needs the highest resolution
over a large area. In most cases the users fall into 2 groups: Users who
normally use the microscope to take images, then use the entire negative to
make a print. These are often biologists, who do not need sub-nm resolution.
In these cases the real issue is field of view and the appropriate camera
would be a side-mount camera.

On the other end are users who need the highest resolution (for example
materials scientists who do lattice imaging). Normally they need the highest
resolution in a very small area, though, and not over the entire field of a
negative. When I used to do this, it often turned out, that the sample was
slightly curved or other artifacts prohibited a good lattice imaging in
areas that were only nms away from the area I was looking at.

For example, if you take images at 500kx (for lattice imaging), the entire
field of view is roughly 200 nm. That is about 1000 lattice spacings in Si.
It is very unlikely, that you can keep the imaging conditions constant over
this distance to allow the use of the full negative.

And if you DO need higher resolution with a larger field of view, you can
always take several images with a slight overlap and montage them
electronically. In our software you can even do that automatically, provided
of course that you have a motorized stage. That way you can get images that
do approach the information density of film. However, the files tend to be
huge and can become a pain to work with.

Now, if we talk about information density, the amount of information per
pixel (or per unit area) has to be taken into account also. Typical CCDs
today are very linear and have 12- 14 bit information depth per pixel. That
corresponds to 4,000 to 16,000 levels of gray. Film, on the other hand has a
very non-linear characteristic, which makes it hard to get quantitative
information out of the film. As to the "bit -depth" of film: Film is really
a binary medium -- a grain can either be exposed or not. As far as I know,
there are no "partially exposed grains". If a grain in the film is about 5
microns in diameter, the film would have a 1-bit information depth on a 5
micron scale (on or off). On a 25 micron scale that would then roughly be a
5-bit resolution. In other words, a digital camera with 24 microns pixel
resolution has a much better bit-depth than film. Of course this is only a
crude approximation, as it does not take into account overlapping grains,
beam spread in the emulsion and other factors, so don't pin me on the
numbers.

mike

} } } } } } } } } } WE HAVE MOVED { { { { { { { { {
please make a note of the new address below

Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Anthony J. Garratt-Reed [mailto:tonygr-at-mit.edu]
Sent: Tuesday, June 04, 2002 7:52 AM
To: Microscopy-at-sparc5.microscopy.com


Alan is correct about the cost of film, etc. We charge our users 75˘US per
sheet for film to cover this cost. I disagree, though, about seeing the
image is in focus - if you have a digital system then you can use it to
check the image is in focus if you like, (and it certainly helps at very
high magnifications), but you can still record, with advantage, on film,
when the image will still be in focus.

The issue, though, goes beyond cost. For some users, digital imaging will
give excellent results, and meet all their needs. For others, the extreme
(by current standards) information density of conventional film is still
essential to get the job done. One has to be careful before deciding to
abandon an existing darkroom facility. Of course, the equation could be
different if a new facility is being designed. We still teach our users to
take all their images on film, for subsequent scanning. We do have digital
capture systems, but they are generally used only by outside users who
haven't learned how to handle film.

In changing from film to digital imaging, it is also important to
understand that significant re-education of users is important. The two
media have different characteristics, which must be understood and
accounted for if best results are to be obtained.

Tony.


At 01:34 PM 6/3/2002 -0500, you wrote:
} Bill
}
} What about the cost of film and chemicals over the lifetime of the
} microscope? As a one off charge your dept chair was happy but the
} additional costs of dealing with film are not small. Also as a totally
} digital materials science lab the main additional benefit is being able to
} see immediately the image and know it is in focus etc without wasting film
} and time.
}
} Alan


** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6478
**




From daemon Tue Jun 4 14:00:20 2002



From: Steve Barlow :      sbarlow-at-sciences.sdsu.edu
Date: Tue, 4 Jun 2002 11:52:52 -0700
Subject: repair for old ISI sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers

My archeology trip to the storeroom uncovered a vintage ISI PS-2
sputter coater unit, which did not immediately start up when we
plugged it in and tried to pump it down. Does anyone know who might
service these babies, or have a Rosetta stone (i.e.,service manual
and schematics) for same?

thanks in advance

Steve
--
Dr. Steven Barlow
EM Facility/Biology Dept.
San Diego State University
5500 Campanile Drive
San Diego CA 92182-4614
phone: (619) 594-4523
fax: (619) 594-5676
email: sbarlow-at-sunstroke.sdsu.edu
http://www.sci.sdsu.edu/emfacility

Chairman, Educational Outreach subcommittee
promoting microscopy instruction and increased access to microscopes
Microscopy Society of America
http://www.msa.microscopy.com/


From daemon Tue Jun 4 15:06:13 2002



From: Anthony J. Garratt-Reed :      tonygr-at-mit.edu
Date: Tue, 04 Jun 2002 15:58:29 -0400
Subject: Re: Ge detector problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have seen a somewhat similar thing on a 12-year-old Si detector which I
know to have poor vacuum. The "hump" was more symmetric and much closer to
the parent peak (guessing, I would say at about 60-70% of the parent's
energy), and would appear particularly after I conditioned the detector,
reducing in height over a period of a day or two following the
conditioning. If it becomes useful, I'm sure I could dig out a sample
spectrum.

I have eliminated the effect by warming the detector (with hot water in the
dewar) and thoroughly pumping it (down to 10-8 Torr) before re-cooling. In
my case, this was easy because the detector is windowless, and is installed
on a UHV microscope (I had to bake the 'scope afterwards!). The full-width
at half-max resolution is still excellent, but I do get significant tailing
(i.e. degraded events), and am considering getting a new crystal, because I
assume I have crystal damage (or contamination?).

Tony.



} We have a Ge detector on our EDS system that has been there for several years.
}
} Our normal procedure is to fill the dewar with LN2 twice a week (Monday or
} Tuesday and then Thursday or Friday). We would like to think the dewar had
} enough capacity to last a full week, and it used to. However, in the
} end-of-the-week rush, I forgot to fill the dewar and came in to a warm
} detector on Monday. It had been filled last Tuesday, so something seems to
} be degrading that we no longer can go even 6 days on a fill.
}
} We have an LN2 alarm, and it had worked and presumably cut off the high
} voltage. We refilled the dewar, gave it an hour or more to cool off, then
} tried the system again. We got a spectrum, but with artifacts.
}
} I tried conditioning the detector (three times over 24 hours) and
} recalibrating the electronics. The strobe peak got back into balance, but
} artifacts still remain. I have posted three documents on the web at
} ftp://www.marl.iastate.edu/Ge_detector/ for those that might be interested
} in examining the problem for themselves. One file is a bitmap of a line
} trace, another is a color bitmap, and the third, most detailed view, is a
} figure embedded in a word document.
}
} All three files show the same thing - an overlay of spectra from an
} aluminum sample holder and a titanium standard. Both show the
} characteristic peak but with perhaps more of a low energy tail. But the
} real problem is a large hump in the background down-scale from the peak.
} The hump follows the characteristic peak up and down the energy scale with
} the element and always seems to peak at about 38% of the characteristic energy.
}
} I have been given one diagnosis and prognosis from the manufacturer. But I
} wonder if the resident experts on this list might have a second opinion. I
} welcome your comments, particularly if they might lead to a quicker,
} cheaper fix than sending the detector back.
}
} Warren
}
} -------------------------------------------
} No files should be attached to this message
} -------------------------------------------
} Warren E. Straszheim, Ph.D.
} Materials Analysis and Research Lab
} Iowa State University
} 23 Town Engineering
} Ames IA, 50011-3232
}
} Ph: 515-294-8187
} FAX: 515-294-4563
}
} E-Mail: wesaia-at-iastate.edu
} Web: www.marl.iastate.edu
}
} Scanning electron microscopy, x-ray analysis, and image analysis of materials
} Computer applications and networking
}
}


** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6478
**




From daemon Tue Jun 4 15:25:08 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Tue, 04 Jun 2002 13:35:40 -0700
Subject: Re: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Donald

Before I bought the digital camera, I was looking on that image plate
technology. It seems to me, it does not have serious advantage for most of
us. I would like to concentrate my findings in a few sentences.

Image plate PRO:
- has more pixels that most digital cameras (some top-end cameras
HAS similar amount).
- has dynamic range same as digital cameras 12-16 bit.
- it's more sensitive than film.
- very good for electron diffraction experiments!!! (only one real plus to me).
- manufacturers claimed that pixel resolution on the plate is comparable
with film (it's difficult to proof because you have to scan film to
compare, so it'll depend from the film scanner).

Image plate CONTRA:
Imitate the film procedure - you have to perform all procedures as for
film, load/upload plates into cassettes (in the dark), change the magazine
(wait for vacuum), load plates into the scanner (in the dark I believe),
wait for scanning - 2 (or more, don't remember, up to 5 at full resolution
I believe) min etc. So, it does not eliminate the dark-room, scanning is
slow and then you have to process/save/print data. Time consuming. You
have all disadvantages the classical film use: plate may be scratched
during loading/uploading, deformed, lost, dropped on floor with valuable
image etc. One plate is $100 I believe. No practical use in most
biological applications I believe. I am not warranty that all my
information is current, I was shopping for image plate system a few years
ago and my comments represent the situation of that time. Sorry,
manufacturing gays. Sergey

At 08:07 AM 6/4/02 -0500, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

------------------------------------------------------

Sergey Ryazantsev, Ph.D.
Electron Microscopy
Department of Biological Chemistry, School of Medicine
University of California, Los Angeles
Box 951737
Los Angeles, CA 90095-1737

(310) 825-1144 (office)
Pager: (310) 845-0248
FAX: (310) 206-5272 (departmental)
mailto:sryazant-at-ucla.edu



From daemon Tue Jun 4 15:48:30 2002



From: Sinkler, Wharton :      WSinkler-at-uop.com
Date: Tue, 4 Jun 2002 15:41:15 -0500
Subject: RE: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sergey et al,

I'd just like to point out that instead of relying on uncertain
recollections, there is some good work in the literature which discusses
both slow-scan CCD's and imaging plates. Here's a start:

J. M. Zuo "Electron detection characteristics of slow-scan CCD camera"
Ultramicroscopy 66 (1996) 21-33.

J. M. Zuo, M. R. McCartney and J. C. H. Spence "Performance of imaging
plates for electron recording" Ultramicroscopy 66 (1996) 35-47

G. Y. Fan and M. H. Ellisman "Digital imaging in transmission electron
microscopy" Journal of Microscopy 200 (2000) 1-13.

Although the technology is continuously evolving, the basic points are still
valid.

Regards,
Wharton

*************************************************
Wharton Sinkler, Ph.D.
UOP LLC
25 E. Algonquin Rd.
Des Plaines, IL 60017-5017
847-391-3878


} -----Original Message-----
} From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} Sent: Tuesday, June 04, 2002 3:36 PM
} To: microscopy-at-sparc5.microscopy.com
} Subject: Re: TEM imaging plate technology
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear Donald
}
} Before I bought the digital camera, I was looking on that image plate
} technology. It seems to me, it does not have serious advantage for most
} of
} us. I would like to concentrate my findings in a few sentences.
}
} Image plate PRO:
} - has more pixels that most digital cameras (some top-end cameras
} HAS similar amount).
} - has dynamic range same as digital cameras 12-16 bit.
} - it's more sensitive than film.
} - very good for electron diffraction experiments!!! (only one real plus to
} me).
} - manufacturers claimed that pixel resolution on the plate is comparable
} with film (it's difficult to proof because you have to scan film to
} compare, so it'll depend from the film scanner).
}
} Image plate CONTRA:
} Imitate the film procedure - you have to perform all procedures as for
} film, load/upload plates into cassettes (in the dark), change the magazine
}
} (wait for vacuum), load plates into the scanner (in the dark I believe),
} wait for scanning - 2 (or more, don't remember, up to 5 at full resolution
}
} I believe) min etc. So, it does not eliminate the dark-room, scanning is
} slow and then you have to process/save/print data. Time consuming. You
} have all disadvantages the classical film use: plate may be scratched
} during loading/uploading, deformed, lost, dropped on floor with valuable
} image etc. One plate is $100 I believe. No practical use in most
} biological applications I believe. I am not warranty that all my
} information is current, I was shopping for image plate system a few years
} ago and my comments represent the situation of that time. Sorry,
} manufacturing gays. Sergey
}
} At 08:07 AM 6/4/02 -0500, you wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} }
} }
} } Dear EM Netters,
} }
} } Does anyoun out there have experience with imaging plate technology (i.e.
} } Ditabis)? I
} } would like to hear pros and cons concidering this technology. How does
} it
} } compare to
} } digital CCD camera systems (cost, time, supplies, reliability. We are
} } looking into possibly
} } retrofitting our JEOL 100CX with an off-line system.
} }
} }
} } Thanks in advance
} }
} }
} } Donald G. Awbrey, HT (ASCP), QIHC
} } Electron Microscopy / Image Analysis
} } 817-878-5647
} } donaldawbrey-at-texashealth.org
} }
}
} ------------------------------------------------------
}
} Sergey Ryazantsev, Ph.D.
} Electron Microscopy
} Department of Biological Chemistry, School of Medicine
} University of California, Los Angeles
} Box 951737
} Los Angeles, CA 90095-1737
}
} (310) 825-1144 (office)
} Pager: (310) 845-0248
} FAX: (310) 206-5272 (departmental)
} mailto:sryazant-at-ucla.edu
}


From daemon Tue Jun 4 18:38:22 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Tue, 4 Jun 2002 19:35:19 -0400
Subject: Microprobe with FEG

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sarath;

I'm sure you will get many opinions on this question but I'll offer mine
anyway even though there are plenty of folks out there that have more
specific knowledge and may provide empirical comparisons and some literature
references. However, I have both a LaB6 and a FESEM side by side, both with
identical Si/sapphire SUTW windows, light element detectors, going into a
common processor from the same vendor. In my opinion, and some considerable
experience in this comparison, unless doing very low voltage work such as
high resolution elemental mapping of light elements,e.g. C, O, F, N, which
is more easily done with a FEG, the physics of electron-solid interactions
essentially dictate the results. My biggest source of error [spatially] is
beam drift when doing elemental mapping and that is influenced, in my
experience, much more by mechanical vibration, sample charging, stage
moving, heavy footed people near the column etc., than beam stability, beam
shape, astigmatism etc.

In any event, a coulomb of charge, is a coulomb of charge, no matter what
gun it originates from. The physics of the electron-solid interaction
doesn't care which gun the electron was pulled out of. But for high mag.,
high res. imaging, I'll take the FEG all the time. The nice thing about
running the lab., AND working in it, is you get to tell everyone else they
have to use the "other" SEM.

Peter Tomic
Failure Analysis & Analytical Services
Anadigics, Inc.

-----Original Message-----
} From: Menon Sarath K Contr AFRL/MLLM [mailto:Sarath.Menon-at-wpafb.af.mil]
Sent: Tuesday, June 04, 2002 11:42 AM
To: 'Microscopy-at-MSA.Microscopy.Com'


Could experts comment on advantages and disadvantages of using a microprobe
with a FEG vs. a LaB6 emitter?
Pointers to any recent publications will be nice too.

thanks

Sarath Menon
Scientist
UES Inc.
Dayton, OH


From daemon Tue Jun 4 19:17:24 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Tue, 4 Jun 2002 20:16:01 -0400
Subject: Re: Philips EM400 Matching Mains Transformer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Eric;

I'm actually surprised that the two legs are within a volt! You must have a
better power company than we do. If you put a strip-chart recorder on your
A.C. line, you will see all kinds of fluctuations over time, especially
during high demand periods in the summer months. In any event, your
instrument should have a voltage regulator within it and a step-down
transformer, often to 100 volts. I wouldn't spend the money on a line
regulator or U.P.S. unless you find some issues with the instrument since
its' power supply should be capable of "cleaning" up a noisy line . Unless
the instrument was designed by a high school student in Wood Shop, the D.C.
supply should be well regulated with little ripple or A.C. component in its'
D.C. supply and generally immune to most A.C. line fluctuations.

We lost power yesterday for 10 seconds and a turbomolecular pump on an FIB
rotated it's last rotation. Now we're talking real misery.

Peter

-----Original Message-----
} From: Eric Anderson [mailto:anderson_e-at-southernct.edu]
Sent: Tuesday, June 04, 2002 12:54 PM
To: Microscopy-at-sparc5.microscopy.com


} Greetings All!
}
} We are in the process of setting up a second-hand EM400 acquired
} recently, and have found that the mains matching transformer was not
} included. Our raw supply main is 237 volts/60Hz (118V on one leg,
} 119V on the other, stability unknown), and I'm thinking this is not
} close enough to the specified 220V to go without the transformer. Any
} ideas? If we do need some line conditioning, can anyone recommend a
} particular device, or source for the original Philips transformer?
}
} Many thanks for any tips!
} -Eric
} --
}
Eric Anderson
SCSU Physics Adjunct
203-392-6455
anderson_e-at-southernct.edu

}



From daemon Tue Jun 4 19:48:30 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Tue, 04 Jun 2002 17:41:40 -0700
Subject: RE: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Wharton

Are you seriously think, people will read that very specific papers? Do
you think, so many of us is ready to go through all this math etc? I agree,
it's very important to understand the scientific foundation for device you
are going to use. But, in practice you have deal not with such nice things
as PSF (point spread function) or MTF. The camera with very nice PSF may
have terrible design and not suitable for use. Company, who manufactured
that baby may not respond on your telephone calls and finally you may
figure out that it does not exists anymore. I do read scientific journals
and it's extremely easy to find informations on the Internet nova days,
ListServer to me is a place where we have chance to share our personal
experience (bad or good), which mostly is not published and not indexed by
Medline/CC whatever. Anyway, thank you for the reference, I enjoyed
reading it. Sergey

At 01:41 PM 6/4/02, you wrote:

} Sergey et al,
}
} I'd just like to point out that instead of relying on uncertain
} recollections, there is some good work in the literature which discusses
} both slow-scan CCD's and imaging plates. Here's a start:
}
} J. M. Zuo "Electron detection characteristics of slow-scan CCD camera"
} Ultramicroscopy 66 (1996) 21-33.
}
} J. M. Zuo, M. R. McCartney and J. C. H. Spence "Performance of imaging
} plates for electron recording" Ultramicroscopy 66 (1996) 35-47
}
} G. Y. Fan and M. H. Ellisman "Digital imaging in transmission electron
} microscopy" Journal of Microscopy 200 (2000) 1-13.
}
} Although the technology is continuously evolving, the basic points are still
} valid.
}
} Regards,
} Wharton
}
} *************************************************
} Wharton Sinkler, Ph.D.
} UOP LLC
} 25 E. Algonquin Rd.
} Des Plaines, IL 60017-5017
} 847-391-3878
}
}
} } -----Original Message-----
} } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } Sent: Tuesday, June 04, 2002 3:36 PM
} } To: microscopy-at-sparc5.microscopy.com
} } Subject: Re: TEM imaging plate technology
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Dear Donald
} }
} } Before I bought the digital camera, I was looking on that image plate
} } technology. It seems to me, it does not have serious advantage for most
} } of
} } us. I would like to concentrate my findings in a few sentences.
} }
} } Image plate PRO:
} } - has more pixels that most digital cameras (some top-end cameras
} } HAS similar amount).
} } - has dynamic range same as digital cameras 12-16 bit.
} } - it's more sensitive than film.
} } - very good for electron diffraction experiments!!! (only one real plus to
} } me).
} } - manufacturers claimed that pixel resolution on the plate is comparable
} } with film (it's difficult to proof because you have to scan film to
} } compare, so it'll depend from the film scanner).
} }
} } Image plate CONTRA:
} } Imitate the film procedure - you have to perform all procedures as for
} } film, load/upload plates into cassettes (in the dark), change the magazine
} }
} } (wait for vacuum), load plates into the scanner (in the dark I believe),
} } wait for scanning - 2 (or more, don't remember, up to 5 at full resolution
} }
} } I believe) min etc. So, it does not eliminate the dark-room, scanning is
} } slow and then you have to process/save/print data. Time consuming. You
} } have all disadvantages the classical film use: plate may be scratched
} } during loading/uploading, deformed, lost, dropped on floor with valuable
} } image etc. One plate is $100 I believe. No practical use in most
} } biological applications I believe. I am not warranty that all my
} } information is current, I was shopping for image plate system a few years
} } ago and my comments represent the situation of that time. Sorry,
} } manufacturing gays. Sergey
} }
} } At 08:07 AM 6/4/02 -0500, you wrote:
} } } ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------.
} } }
} } }
} } }
} } }
} } } Dear EM Netters,
} } }
} } } Does anyoun out there have experience with imaging plate technology (i.e.
} } } Ditabis)? I
} } } would like to hear pros and cons concidering this technology. How does
} } it
} } } compare to
} } } digital CCD camera systems (cost, time, supplies, reliability. We are
} } } looking into possibly
} } } retrofitting our JEOL 100CX with an off-line system.
} } }
} } }
} } } Thanks in advance
} } }
} } }
} } } Donald G. Awbrey, HT (ASCP), QIHC
} } } Electron Microscopy / Image Analysis
} } } 817-878-5647
} } } donaldawbrey-at-texashealth.org
} } }
} }
} } ------------------------------------------------------
} }
} } Sergey Ryazantsev, Ph.D.
} } Electron Microscopy
} } Department of Biological Chemistry, School of Medicine
} } University of California, Los Angeles
} } Box 951737
} } Los Angeles, CA 90095-1737
} }
} } (310) 825-1144 (office)
} } Pager: (310) 845-0248
} } FAX: (310) 206-5272 (departmental)
} } mailto:sryazant-at-ucla.edu
} }

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Wed Jun 5 00:30:18 2002



From: Bill Miller :      microbill-at-mohawk.net
Date: Wed, 05 Jun 2002 05:53:47 -0400
Subject: Re: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Eric,

237V is a bit too high. It must be between 208V and 220V. The simplest and
economical way to power up your TEM is to use Buck and Boost transformer
connected as autotransformer. Philips matching unit originally supplied with
EM400, was also an autotransformer. Order from Grainger, www.grainger.com ,
many locations everywhere, stock # 1H270, $174. This one will reduce your
line voltage by 24V, making it 213V to 211V (see below), which is perfect.

Notes.
1) Make sure that grounding is correct and safe.
2)This unit can be connected in 4 different ways, for increasing or reducing
the line voltage by 12V or 24V, and is shipped with 1 page manual. Read it.
3) The above part number is given for your particular (line voltage) case.
Others may require a transformer with different part number.
4) Transformer will not stabilize the line voltage, only change it.
5) I assume that you measured line voltage with no load connected- the
actual voltage may drop about 1V or 2V when you connect the TEM.

Vitaly Feingold
Scientific Instruments and Applications
2773 Heath Lane, Duluth GA 30096
(770)232-7785 ph.
(770)232-1791 fax
(678)467-0012 mobile

This message is made of 100% recycled electrons.
----- Original Message -----
} From: Eric Anderson {anderson_e-at-southernct.edu}
To: {Microscopy-at-sparc5.microscopy.com}
Sent: Tuesday, June 04, 2002 12:53 PM


The Ditabis plates now offer with resolution up to 6000x5000 and 20 bit
data for FAR less money than a lower resolution CCD solution. The newest
system even offers variable resolution. The higher sensitivity is often
useful for low dose images as well. There are applications where one or the
other may be the right choice but certainly biological image should not be
excluded. You failed to mention any of the downside for CCD devices - of
which there is, of course, a comparable list that includes limited image
area and dynamic range, image resolution and cost.

Bill Miller
ElectroImage


At 01:35 PM 6/4/2002 -0700, Sergey Ryazantsev wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Wed Jun 5 05:49:10 2002



From: Khalid Boulahya :      khalid-at-quim.ucm.es
Date: Wed, 05 Jun 2002 12:42:40 +0200
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




**************************************************************
Khalid Boulahya
Depto de Química Inorgánica
Facultad De Químicas
Universidad Complutense de Madrid (UCM)
28040 Spain
khalid-at-quim.ucm.es
**************************************************************



From daemon Wed Jun 5 05:50:23 2002



From: Khalid Boulahya :      khalid-at-quim.ucm.es
Date: Wed, 05 Jun 2002 12:45:04 +0200
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe

**************************************************************
Khalid Boulahya
Depto de Química Inorgánica
Facultad De Químicas
Universidad Complutense de Madrid (UCM)
28040 Spain
khalid-at-quim.ucm.es
**************************************************************



From daemon Wed Jun 5 06:31:42 2002



From: Khalid Boulahya :      khalid-at-quim.ucm.es
Date: Wed, 05 Jun 2002 13:24:48 +0200
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe



From daemon Wed Jun 5 07:19:06 2002



From: Ingo Daberkow :      ingo.daberkow-at-tvips.com
Date: Wed, 05 Jun 2002 14:10:51 +0200
Subject: Re: TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Donna,

} From the camera resolution point of view, in principle, there is no difference
between the side-mounted and bottom-mounted position. Of course the
electron-optical resolution has to be adapted accordingly – in order to acquire
the wished field of view. For example, the mag must be increased by of factor of
3x if a side-mounted cameras is used. This is not easy possible for high
resolution work, therefore mainly bottom-mounted cameras are used for such
applications.

But there is an essential difference between both positions: The used aperture
of the bottom mounted camera is much smaller, this means the image is captured
very close to the optical axis. Therefore these images will show almost no
distortions which are due to the aberrations of the projective lenses.
At the side-mounted position these aberrations can be up to 1-2% which seems not
to be very much. But for a 1k camera, this means a displacement of 10-20 pixels
which makes problems when images are stitched together.
In practise, if you consider to combine images together (stitching, tiling) - in
order to get a higher resolution or larger field of view - you should consider a
bottom-mounted camera.

I hope this helps to clarify the situation.

Best wishes,
Ingo

+++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
We have moved. Please notice our new address.

Dr. Ingo Daberkow
Tietz Video and Image Processing Systems GmbH
Eremitenweg 1
D-82131 Gauting, Germany
Tel: +49-89-8506567
FAX: +49-89-8509488
Internet: http://www.tvips.com/
Email: ingo.daberkow-at-tvips.com
+++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++



Clarkson Donna R Contr USAMRD wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Chuck,
} Thank-you for your input. As with many others the opinions are mixed,
} but seem to lean toward the AMT camera system. I am a little confused on one
} issue, though. You mentioned that you have lost some resolution by
} side-mounting the camera; others have noted that they thought they had lost
} resolution by under-the-column mounting. Hmmmm, I guess it depends on the
} scope it's mounted to.
}
} Best regards,
}
} Donna
}
} Northrop Grumman Information Technology
} for U S Army Medical Research Detachment
} at Brooks Air Force Base
} Phone (210) 536-1416
} FAX (210) 536-1449
} e-mail donna.clarkson-at-brooks.af.mil
}
} -----Original Message-----
} } From: Chuck Butterick [mailto:cbutte-at-ameripol.com]
} Sent: Tuesday, June 04, 2002 8:15 AM
} To: Microscopy-at-sparc5.microscopy.com; donna.clarkson-at-brooks.af.mil
} Subject: Re: TEM-Digital Camera Systems
}
} We upgraded an old Philips 300 with an AMT system with a Hamamatsu
} C4742-95 camera 3.5 years ago. The system was mounted in the old 35mm
} camera port, which does not have the resolution an under-the-column
} camera will have, because I didn't want to lose the capability of
} using film. The system has allowed a superior throughput of work with
} same day turnarounds on projects that took 2 weeks before. Further,
} digital image acquisition permits image analysis without any
} intervening steps, which is crucial, for example, to calculating the
} particle size of carbon black aggregates in a time-conscious
} production environment.
}
} For every day work, a laser printer works just fine. When printing
} images for show/publication, an inkjet printer using a high quality
} inkjet paper (Kodak or HP photo quality papers are my choice) does
} very well. The ability to make high quality enlargements, though, is
} restricted. Wet chemistry, film and paper is better, but not by
} much...and, arguably, may not be worth the extra time and effort.
}
} Not only does a digital system save on film, paper and chemistry (with
} all the associated environmental concerns), it saves on
} labor/time...and that is where the largest dollar savings is. Further,
} as happened in my situation, I didn't need a darkroom any more so that
} room was converted to house the new SEM/EDX system, which is a more
} profitable use of the square footage. Even academic chairman (well,
} in medical schools anyway) are concerned about the amount of research
} dollars per square foot.
}
} The choice of a system must always factor in service as well as
} capability and price. AMT and Gatan both have excellent systems.
} However, no one, given my circumstances at the time, would have chosen
} an alternate system. Given another chance to buy a digital system for
} a TEM, the AMT folks would have first shot, with the alternates having
} an uphill, but not impossible, battle to convince me otherwise.
}
} AMT competitors make excellent products. I'm just very satisfied with
} cost, service, and instrumentation obtained from AMT. The usual
} disclaimers apply.
}
} Chuck Butterick
} Degussa Corporation
} Borger, TX
}
} ______________________________ Reply Separator
} _________________________________
} Subject: TEM-Digital Camera Systems
} Author: Clarkson Donna R Contr USAMRD {donna.clarkson-at-brooks.af.mil} at
} INTERNET-MAIL
} Date: 5/31/02 2:33 PM
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} We are looking to purchase a digital camera system for our TEM and
} would like some feedback from those of you who already utilize such systems.
}
} Some questions I'd like to ask are:
}
} * What make & model camera system do you have?
} * What type, make & model printer(s) do you use, (e.g. inkjet,
} dye-sublimation, silver halide)?
} * Do you feel you get micrograph-quality resolution and images
} utilizing that system?
} * Ease of use--is it user-friendly or cumbersome?
} * Ease and/or quality of service--are any of the components
} serviceable by you? How quickly & easily is it to get a service rep?
} * How long have you had this system?
} * Would you recommend this system--why or why not?
} * What, if anything, might you have done differently?
}
} I'd greatly appreciate any responses. Thank-you in advance for your
} help.
}
} Donna R. Clarkson
}
}



From daemon Wed Jun 5 07:51:30 2002



From: Paula Sicurello :      patpxs-at-gwumc.edu
Date: Wed, 05 Jun 2002 08:43:53 -0400
Subject: Re: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I was just dealing with the company that will be the US rep. for Ditabis. They are ElectroImage at www.electroimage.com
As far as I can tell, the plates are not light sensitive until after they have been exposed and each plate is useful for up to 1000 exposures (you just "erase" the previous image). Iwas told that the unit with all the bells and whistles will cost around $50 to $60K. I think it might be a good thing for people who want digital, but I never saw it in action.

We still are old fashioned and want negatives that we can archive.

Paula :-)

p.s. I have no financial interest in ElectroImage I just know that they are going to sell the beasty.

Paula Sicurello
George Washington Univ. Medical Center
Dept. of Pathology, Ross Hall rm 505
Electron Microscope Lab
2300 Eye St.
Washington, DC 20037
202-994-2930 phone
202-994-2518 fax

} } } "Awbrey, Donald" {DonaldAwbrey-at-texashealth.org} 06/04/02 09:07AM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America




Dear EM Netters,

Does anyoun out there have experience with imaging plate technology (i.e.
Ditabis)? I
would like to hear pros and cons concidering this technology. How does it
compare to
digital CCD camera systems (cost, time, supplies, reliability. We are
looking into possibly
retrofitting our JEOL 100CX with an off-line system.


Thanks in advance


Donald G. Awbrey, HT (ASCP), QIHC
Electron Microscopy / Image Analysis
817-878-5647
donaldawbrey-at-texashealth.org





From daemon Wed Jun 5 08:31:08 2002



From: Sinkler, Wharton :      WSinkler-at-uop.com
Date: Wed, 5 Jun 2002 08:23:15 -0500
Subject: RE: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



OK, Truce Sergey!

Your point is well taken, and yes, design considerations and convenience may
overwhelm a good MTF or linearity.

However, I think you are mistaken about dynamic range being similar for
Imaging Plates and CCD's. That being said, I have gone back to look at the
papers I mentioned and I'm not able to find a clear statement of what I
mean.

I believe (maybe someone will set us straight on this) that the saturation
of the IP is considerably higher than the CCD. By 'saturation' I mean the
dose (electrons/unit area) at which there is significant deviation from
linearity.

The reason saturation level is important (and why the IP is great for
diffraction) is that signal/noise at low electron doses is dominated by shot
noise. So if you are trying to record a diffraction pattern with a CCD with
a single exposure without saturating the CCD at the strong spots, the weak
reflections will be noisy simply because relatively few electrons are
arriving there (statistical or 'shot' noise). With a higher saturation (I
believe the case for IP), you can expose longer and get better statistics
overall. (Of course you can also record multiple exposures with the CCD at
different times and merge the data).

Regards,
Wharton



} -----Original Message-----
} From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} Sent: Tuesday, June 04, 2002 7:42 PM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: RE: TEM imaging plate technology
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear Wharton
}
} Are you seriously think, people will read that very specific papers? Do
} you think, so many of us is ready to go through all this math etc? I
} agree,
} it's very important to understand the scientific foundation for device you
}
} are going to use. But, in practice you have deal not with such nice
} things
} as PSF (point spread function) or MTF. The camera with very nice PSF may
} have terrible design and not suitable for use. Company, who manufactured
} that baby may not respond on your telephone calls and finally you may
} figure out that it does not exists anymore. I do read scientific
} journals
} and it's extremely easy to find informations on the Internet nova days,
} ListServer to me is a place where we have chance to share our personal
} experience (bad or good), which mostly is not published and not indexed by
}
} Medline/CC whatever. Anyway, thank you for the reference, I enjoyed
} reading it. Sergey
}
} At 01:41 PM 6/4/02, you wrote:
}
} } Sergey et al,
} }
} } I'd just like to point out that instead of relying on uncertain
} } recollections, there is some good work in the literature which discusses
} } both slow-scan CCD's and imaging plates. Here's a start:
} }
} } J. M. Zuo "Electron detection characteristics of slow-scan CCD camera"
} } Ultramicroscopy 66 (1996) 21-33.
} }
} } J. M. Zuo, M. R. McCartney and J. C. H. Spence "Performance of imaging
} } plates for electron recording" Ultramicroscopy 66 (1996) 35-47
} }
} } G. Y. Fan and M. H. Ellisman "Digital imaging in transmission electron
} } microscopy" Journal of Microscopy 200 (2000) 1-13.
} }
} } Although the technology is continuously evolving, the basic points are
} still
} } valid.
} }
} } Regards,
} } Wharton
} }
} } *************************************************
} } Wharton Sinkler, Ph.D.
} } UOP LLC
} } 25 E. Algonquin Rd.
} } Des Plaines, IL 60017-5017
} } 847-391-3878
} }
} }
} } } -----Original Message-----
} } } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } } Sent: Tuesday, June 04, 2002 3:36 PM
} } } To: microscopy-at-sparc5.microscopy.com
} } } Subject: Re: TEM imaging plate technology
} } }
} } }
} ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} -----------------------------------------------------------------------.
} } }
} } }
} } } Dear Donald
} } }
} } } Before I bought the digital camera, I was looking on that image plate
} } } technology. It seems to me, it does not have serious advantage for
} most
} } } of
} } } us. I would like to concentrate my findings in a few sentences.
} } }
} } } Image plate PRO:
} } } - has more pixels that most digital cameras (some top-end cameras
} } } HAS similar amount).
} } } - has dynamic range same as digital cameras 12-16 bit.
} } } - it's more sensitive than film.
} } } - very good for electron diffraction experiments!!! (only one real
} plus to
} } } me).
} } } - manufacturers claimed that pixel resolution on the plate is
} comparable
} } } with film (it's difficult to proof because you have to scan film to
} } } compare, so it'll depend from the film scanner).
} } }
} } } Image plate CONTRA:
} } } Imitate the film procedure - you have to perform all procedures as for
} } } film, load/upload plates into cassettes (in the dark), change the
} magazine
} } }
} } } (wait for vacuum), load plates into the scanner (in the dark I
} believe),
} } } wait for scanning - 2 (or more, don't remember, up to 5 at full
} resolution
} } }
} } } I believe) min etc. So, it does not eliminate the dark-room, scanning
} is
} } } slow and then you have to process/save/print data. Time consuming.
} You
} } } have all disadvantages the classical film use: plate may be scratched
} } } during loading/uploading, deformed, lost, dropped on floor with
} valuable
} } } image etc. One plate is $100 I believe. No practical use in most
} } } biological applications I believe. I am not warranty that all my
} } } information is current, I was shopping for image plate system a few
} years
} } } ago and my comments represent the situation of that time. Sorry,
} } } manufacturing gays. Sergey
} } }
} } } At 08:07 AM 6/4/02 -0500, you wrote:
} } }
} } ------------------------------------------------------------------------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } } } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } } } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} } -----------------------------------------------------------------------.
} } } }
} } } }
} } } }
} } } }
} } } } Dear EM Netters,
} } } }
} } } } Does anyoun out there have experience with imaging plate technology
} (i.e.
} } } } Ditabis)? I
} } } } would like to hear pros and cons concidering this technology. How
} does
} } } it
} } } } compare to
} } } } digital CCD camera systems (cost, time, supplies, reliability. We
} are
} } } } looking into possibly
} } } } retrofitting our JEOL 100CX with an off-line system.
} } } }
} } } }
} } } } Thanks in advance
} } } }
} } } }
} } } } Donald G. Awbrey, HT (ASCP), QIHC
} } } } Electron Microscopy / Image Analysis
} } } } 817-878-5647
} } } } donaldawbrey-at-texashealth.org
} } } }
} } }
} } } ------------------------------------------------------
} } }
} } } Sergey Ryazantsev, Ph.D.
} } } Electron Microscopy
} } } Department of Biological Chemistry, School of Medicine
} } } University of California, Los Angeles
} } } Box 951737
} } } Los Angeles, CA 90095-1737
} } }
} } } (310) 825-1144 (office)
} } } Pager: (310) 845-0248
} } } FAX: (310) 206-5272 (departmental)
} } } mailto:sryazant-at-ucla.edu
} } }
}
} _____________________________________
}
} Sergey Ryazantsev Ph. D.
} Electron Microscopy
} UCLA School of Medicine
} Department of Biological Chemistry
} Box 951737
} Los Angeles, CA 90095-1737
}
} Phone: (310) 825-1144
} FAX (departmental): (310) 206-5272
} mailto:sryazant-at-ucla.edu
}
}
}


From daemon Wed Jun 5 08:31:31 2002



From: Clarkson Donna R Contr USAMRD :      donna.clarkson-at-brooks.af.mil
Date: Wed, 5 Jun 2002 08:25:49 -0500
Subject: Re:TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sergey,
Don't worry about sounding like an advertisement. The more information
I can gather on the digital cameras the better. I did find it rather amusing
to read your "disclaimer" at the end stating that you were not bribed with
dinner or coffee! Good sense of humor! Thanks.

Donna R. Clarkson




-----Original Message-----
} From: Sergey Ryazantsev [mailto:sryazant-at-ucla.edu]
Sent: Monday, June 03, 2002 11:02 PM
To: Microscopy-at-sparc5.microscopy.com


Using digital camera (600W BioScan, Gatan) with TEM over last couple of
month I find the following.

-most people prefer to use digital imaging even we offer equal access to
the film as well. It's because in most cases people need to evaluate
sample quality etc, they simply don't need film quality. From another hand
they could use film anytime if they wanted.

-with digital images you have chance to see your sample immediately
(important for evaluation/students in rush/deadlines etc).

-If you do not enlarge the digital image, 7x7" draft printout on laser
printer (standard) is looks very similar to the image produced from scanned
film. Manipulating with image resolution, dpi etc, you should keep in
mind, that 1200 dpi printer resolution mean that printer could produce 1200
dots per inch in the row making solid line. In order to create shadows of
gray, printer should print dots with space. For instance, 50% of grey
would be represented by 'dot-space' sequence. It mean, that resolution
would be 50% from 1200 = 600 dpi. For 30% gray, it would be 400 dpi. So,
there is no reason to abuse printer sending 1200 dpi 50 Mb image. On
practice, I could not recognize difference between 1x1K at 144 dpi digital
camera image from scanned from film 1200 dpi image: On laserJet (similarly
on our Tektronix dye-sub) - they are looks very the same. BUT: the
difference is that you could enlarge 1200 dpi image (and print with the
same quality) and you COULDN'T do so with 1x1K image from digital
camera. Another remark here: in 'prestigious' magazines like
Science/Nature technical editors have tendency to reduce image to the
postal stamp size, so you could do it by yourself and there is no need to
use film for such small images, digital camera will work just fine!

-another advantage of the digital camera, which, actually, I did not
expect, is its sensitivity. My camera is at least x10 more sensitive than
film, so exposure time is 0.1-0.2 sec versus my usual 1.5-2 sec for the
film, So, less drift and sample damage.

-another things are educational. Instead keep students in the dark and
teach them how to adjust binoculars (instead doing actual EM), we are
comfortably sitting around the 19" flat screen monitor with lights ON (and
nice, not loud classic music). I find it's more convenient to show how to
focus image on the computer screen rather than make a 'focus series', then
develop/print it. At this point students usually don't remember how image
was looks like on the microscope's screen and we have to start again. I
don't mean that students don't need to know how to operate
microscope/binoculars, I just mean that digital camera makes this process
more enjoyable and creative.

-another thing is academic: The live image from digital camera I broadcast
to the Departmental network, so you could see live image on any
departmental computer. PIs now comfortably sit in their offices and enjoy
watching how their students working on the TEM. Similarly it works for
Internet. We did a few session when our collaborators observe their
samples sitting far-far away from UCLA.

So, the bottom line is that TEM digital camera is very convenient
'supplement' to your EM. It does not replace the film, but enhance your
microscope's abilities. It's like power-steering in the car: convenient,
but not necessary - you may park your car in NY without this 'supplement'
right? As a matter of fact, cars without power-steering is cheaper.

Sergey

P.S. I just figured out that this message may be recognized as an
'advertisement'. So, it looks like I was working hard writing on ESL for
manufacturers. I would like to declare openly - I did not intend to do so
and nobody from Gatan or other manufacturer offered to me dinner or coffee
when I was writing it. I did it in sincere believe of what I was writing.








From daemon Wed Jun 5 09:24:33 2002



From: John F. Mansfield :      jfmjfm-at-umich.edu
Date: Wed, 05 Jun 2002 09:51:07 -0400
Subject: Temporary Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello there, the University of Michigan North Campus EMAL has an opening
available for a temporary Research Associate II Engineering. The generic
job description for a Research Associate II Engineering may be found on the
U of M web site (http://www.umich.edu/%7Ehraa/erc/descriptions/cd12871.htm)
The specific requirements for this position are as follows:

Research Assoc. II Engineering
Operate, assist and train users in the use of the following NC EMAL lab
instruments: JEOL 2010 FEG Analytical Transmission Electron Microscope, JEOL
4000EX High Resolution Transmission Electron Microscope, Philips XL30 FEG
SEM Scanning Electron Microscope, ElectroScan E3 Environmental Scanning
Electron Microscope, Digital Instruments Scanning Force Microscope, PHI 5400
X-ray Photo-electron Spectrometer, PHI 660 Scanning Auger Microprobe,
Darkroom facilities, specimen preparation equipment. Assist users with
reduction and analysis of data recorded on same instruments.

Maintenance of some lab equipment including PHI 5400 XPS, PHI 660 Auger,
Environmental SEM, darkroom and specimen preparation equipment. Prepare
internal and external billing for EMAL instrument use each month. Purchase
lab supplies. Take part in collaborative projects with other members of the
university. Develop personal research projects as time permits. Other duties
as defined by supervisor.

The position is full time, but will be short-term only ( less than 11
months)
This position lies in the University Professional/Administrative salary
grade 9 which currently extends from $11.25 per hour to $29.45 per hour
maximum. The actual compensation will depend on the candidate experience.

Minimum qualifications are a bachelors degree in science or engineering and
several years experience in the field of electron microscopy and surface
analysis. Ideal qualifications are a bachelors degree in materials science,
physics, chemistry, engineering or a closely related subject and 5 years of
experience in electron microscopy and surface analysis or a masters degree
in a related subject and 3 years experience in electron microscopy and
surface analysis. Computer literacy is required on two of the following
platforms: PC, Mac, Unix.

Please respond by email or phone to John Mansfield, information in my
signature below.


--
John Mansfield PhD MInstP
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
USA
Phone: (734) 936-3352 FAX (734) 763-2282 Cell. Phone: (734) 834-3913
(Leaving a phone message at 936-3352 is preferable to 834-3913)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html
Location: Lat. 42! 16' 48" Long. 83! 43' 48"



From daemon Wed Jun 5 09:42:29 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Wed, 5 Jun 2002 10:17:44 -0400
Subject: Jeol 100C

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a JEOL 100C top entry scope that is looking for a new home
for anyone interested. It has 2 sets of plate film boxes and holders.
Asking $10,000 or BO with removal and shipping at labor and cost. Scope
has been under service contract with Jeol USA for the last 10 years.


Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University
609-258-5432
jgoodhouse-at-molbio.princeton.edu



From daemon Wed Jun 5 10:03:45 2002



From: Beth Richardson :      beth-at-dogwood.botany.uga.edu
Date: Wed, 5 Jun 2002 10:55:29 -0700
Subject: heating lead citrate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi everyone,
What safety precautions should be followed when heating lead citrate?
I want to try Hanaichi et al's method - A Stable Lead by Modification of
Sato's Method.(1986) J. Electron Microsc. Vol. 35 No. 3, 304-306.
It calls for heating lead citrate for several hours in a melting pot at 200
- 300 C. I have access to a muffle furnace (it's in a hood). Are there
other safety precautions we should use?
thanks in advance for the advice,
Beth

**********************************************************************
Beth Richardson
EM Lab Coordinator
Plant Biology Department
University of Georgia
Athens, GA 30602-7271

Phone - (706) 542-1790 & FAX - (706) 542-1805

"Between the two evils,
I always pick the one I never tried before". Mae West (1893-1980)
**********************************************************************

"And it's only the giving that makes you what you are".
Wond'ring Aloud, Jethro Tull (Aqualung)

***************************************************************************




From daemon Wed Jun 5 10:14:17 2002



From: Mike Marko :      marko-at-wadsworth.org
Date: Wed, 5 Jun 2002 10:59:20 -0400
Subject: TEM - Tomography Workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Colleagues,

A workshop on electron tomography will be held on 14-16 November 2002
in Albany, New York. The workshop is being offered at the Wadsworth
Center laboratories by the Resource for Visualization of Biological
Complexity (RVBC), a Biotechnology Resource supported by NCRR/NIH.

The workshop is intended to provide hands-on experience in electron
tomography, with an emphasis on data collection. Several different data
collection procedures, software packages, and electron microscopes will
be used. There will be daily practical sessions. There will also be
lectures, demonstrations, and practical sessions covering alignment and
reconstruction, visualization techniques, and preparation of
frozen-hydrated specimens.

Instructors and lecturers will include RVBC staff, as well the
individuals directly involved in software and technique development for
the systems that will be used during the workshop.

There is no registration fee.

Registration deadline: 1 October, 2002.

For more information and to register, please go to:

http://home.nycap.rr.com/cdmms/workshop

Michael Marko
Workshop Coordinator
Wadsworth Center, Albany, NY
marko-at-wadsworth.org


From daemon Wed Jun 5 10:31:06 2002



From: Bruce Girrell :      bigirrell-at-microlinetc.com
Date: Wed, 5 Jun 2002 11:25:58 -0400
Subject: RE: repair for old ISI sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I also have an ISI PS-2 coater. When I was looking for documentation, I was
told (by a person who provides maintenance for EM equipment) that the
Polaron E5000 was pretty much the same unit. I got an E5000 manual and the
control descriptions match what I see on my unit, so it seems to be a good
match. I have not opened the unit to compare the schematic against the
actual circuitry, though.

Hope that helps.

Bruce Girrell



From daemon Wed Jun 5 10:31:24 2002



From: Eric Anderson :      anderson_e-at-southernct.edu
Date: Wed, 05 Jun 2002 11:25:22 -0400
Subject: Re: Philips EM400 Matching Mains Transformer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to all for your replies and generous consideration! I think we have
this one problem licked, but our troubleshooting has only just begun. Well,
back to work!

Best regards,
-Eric
---------------------------
Vitaly Feingold wrote:

} Eric,
}
} 237V is a bit too high. It must be between 208V and 220V. The simplest and
} economical way to power up your TEM is to use Buck and Boost transformer
} connected as autotransformer. Philips matching unit originally supplied with
} EM400, was also an autotransformer. Order from Grainger, www.grainger.com ,
} many locations everywhere, stock # 1H270, $174. This one will reduce your
} line voltage by 24V, making it 213V to 211V (see below), which is perfect.
}
} Notes.
} 1) Make sure that grounding is correct and safe.
} 2)This unit can be connected in 4 different ways, for increasing or reducing
} the line voltage by 12V or 24V, and is shipped with 1 page manual. Read it.
} 3) The above part number is given for your particular (line voltage) case.
} Others may require a transformer with different part number.
} 4) Transformer will not stabilize the line voltage, only change it.
} 5) I assume that you measured line voltage with no load connected- the
} actual voltage may drop about 1V or 2V when you connect the TEM.
}
} Vitaly Feingold
} Scientific Instruments and Applications
} 2773 Heath Lane, Duluth GA 30096
} (770)232-7785 ph.
} (770)232-1791 fax
} (678)467-0012 mobile
}
} This message is made of 100% recycled electrons.
} ----- Original Message -----
} From: Eric Anderson {anderson_e-at-southernct.edu}
} To: {Microscopy-at-sparc5.microscopy.com}
} Sent: Tuesday, June 04, 2002 12:53 PM
} Subject: Re: Philips EM400 Matching Mains Transformer
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } } Greetings All!
} } }
} } } We are in the process of setting up a second-hand EM400 acquired
} } } recently, and have found that the mains matching transformer was not
} } } included. Our raw supply main is 237 volts/60Hz (118V on one leg,
} } } 119V on the other, stability unknown), and I'm thinking this is not
} } } close enough to the specified 220V to go without the transformer. Any
} } } ideas? If we do need some line conditioning, can anyone recommend a
} } } particular device, or source for the original Philips transformer?
} } }
} } } Many thanks for any tips!
} } } -Eric
} } } --
} } }
} } Eric Anderson
} } SCSU Physics Adjunct
} } 203-392-6455
} } anderson_e-at-southernct.edu



From daemon Wed Jun 5 11:48:13 2002



From: Stephenson, Matthew :      stephenson-at-impactanalytical.com
Date: Wed, 5 Jun 2002 12:39:41 -0400
Subject: AFM calibration - standards and procedures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings all!

I'm in the process of resuscitating a TopoMetrix TMX 2000 with Discoverer
stage for contact and non-contact AFM. We use three scanners: a 70 micron
tripod, a 7 micron tube, and a 0.76 micron tube. The instrument is up and
running, so the time has come to calibrate, incorporate into our quality
program, and get some customers. My questions are as follows:

1. What reference standards are recommended for x, y, and z distance
calibration? Is anything available that is NIST traceable, or barring that,
certifiable by a reputable authority?

2. Have widely accepted AFM calibration procedures emerged in the
microscopy community? I'd be particularly interested in citations for sound
calibration procedure, and hoary wisdom regarding the effect of instrument
variables on the calibration process and accuracy.

I look forward to your comments!

Matthew K. Stephenson
Analytical Associate
Impact Analytical
1910 West Saint Andrews Road
Midland, MI 48640
(989) 832-5555 X506
stephenson-at-impactanalytical.com




From daemon Wed Jun 5 13:38:14 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Wed, 5 Jun 2002 12:32:26 -0600
Subject: RE: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Careful: the following is an opinion of a CCD system manufacturer. Reader
discretion is advised!

If I remember correctly, the manufacturers for he imaging plates claim a
dynamic range of 20 bit, which is considerably more than the 12 or 14 bit
from a CCD camera. If that, however, makes a big difference for diffraction
patterns I don't know. If you want to record the weakest spots, you need to
expose at least for a time so that the signal is above noise. Since CCD
cameras are very sensitive (down to single electron sensitivity), there is
not much that can be improved by the imaging plates. There are no half
electrons (and if you see 1/3 electrons, book a flight to Stockholm). If the
20 bits are sufficient to image the weakest beams together with the
transmitted beam without saturation I don't know. With the Imaging plates
you can probably get the weakest spots and more medium intensity spots at
the same time than with a CCD. With a CCD you have the advantage that you
immediately see the result and can adjust your exposure accordingly. It's
also very easy to take exposures at different exposure times and combine
them, as the CCDs are very linear.

I can't remember anything about linearity of the imaging plates. But the way
they work (excite crystals into a semi-stable state through the light from a
phosphor and then "read' them by relaxing into ground stage through a
laser), I would think there are some non-linearities as with increasing
exposure you have less and less crystallites that can be excited. Whether
that's critical for quantitative measurements I can't say.

Finally, the images on the plates are stable for a few hours. So you can't
leave the plates in for a few days if you want the best results.

One area, where in my experience (as a TEM user) CCD cameras have a big
advantage is dark field imaging. I remember taking several films with
different exposures to make sure that I had one that was right. With a CCD
there is no guesswork anymore, as you can see the images immediately.

Resolution: theoretically the plate film should have a worse resolution that
film. The electron beam has to travel through the crystallite layer, create
light in the phosphor layer, and the photons then have to travel back to the
crystallites to excite them. This should add to the dispersion of the beam,
resulting in a reduced resolution. The manufacturers of the imaging plates
claim around 20 microns resolution. If that is a 20 micron resolution at 20
bit dynamic range they don't say.

Again, we produce those CCD systems, and I definitely have a bias. But I
thought I put my 2 cents in.

Mike

} } } } } } } } } } WE HAVE MOVED { { { { { { { { {
please make a note of the new address below

Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Sinkler, Wharton [mailto:WSinkler-at-uop.com]
Sent: Wednesday, June 05, 2002 7:23 AM
To: 'Sergey Ryazantsev'; Microscopy-at-sparc5.microscopy.com



OK, Truce Sergey!

Your point is well taken, and yes, design considerations and convenience may
overwhelm a good MTF or linearity.

However, I think you are mistaken about dynamic range being similar for
Imaging Plates and CCD's. That being said, I have gone back to look at the
papers I mentioned and I'm not able to find a clear statement of what I
mean.

I believe (maybe someone will set us straight on this) that the saturation
of the IP is considerably higher than the CCD. By 'saturation' I mean the
dose (electrons/unit area) at which there is significant deviation from
linearity.

The reason saturation level is important (and why the IP is great for
diffraction) is that signal/noise at low electron doses is dominated by shot
noise. So if you are trying to record a diffraction pattern with a CCD with
a single exposure without saturating the CCD at the strong spots, the weak
reflections will be noisy simply because relatively few electrons are
arriving there (statistical or 'shot' noise). With a higher saturation (I
believe the case for IP), you can expose longer and get better statistics
overall. (Of course you can also record multiple exposures with the CCD at
different times and merge the data).

Regards,
Wharton



} -----Original Message-----
} From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} Sent: Tuesday, June 04, 2002 7:42 PM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: RE: TEM imaging plate technology
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear Wharton
}
} Are you seriously think, people will read that very specific papers? Do
} you think, so many of us is ready to go through all this math etc? I
} agree,
} it's very important to understand the scientific foundation for device you
}
} are going to use. But, in practice you have deal not with such nice
} things
} as PSF (point spread function) or MTF. The camera with very nice PSF may
} have terrible design and not suitable for use. Company, who manufactured
} that baby may not respond on your telephone calls and finally you may
} figure out that it does not exists anymore. I do read scientific
} journals
} and it's extremely easy to find informations on the Internet nova days,
} ListServer to me is a place where we have chance to share our personal
} experience (bad or good), which mostly is not published and not indexed by
}
} Medline/CC whatever. Anyway, thank you for the reference, I enjoyed
} reading it. Sergey
}
} At 01:41 PM 6/4/02, you wrote:
}
} } Sergey et al,
} }
} } I'd just like to point out that instead of relying on uncertain
} } recollections, there is some good work in the literature which discusses
} } both slow-scan CCD's and imaging plates. Here's a start:
} }
} } J. M. Zuo "Electron detection characteristics of slow-scan CCD camera"
} } Ultramicroscopy 66 (1996) 21-33.
} }
} } J. M. Zuo, M. R. McCartney and J. C. H. Spence "Performance of imaging
} } plates for electron recording" Ultramicroscopy 66 (1996) 35-47
} }
} } G. Y. Fan and M. H. Ellisman "Digital imaging in transmission electron
} } microscopy" Journal of Microscopy 200 (2000) 1-13.
} }
} } Although the technology is continuously evolving, the basic points are
} still
} } valid.
} }
} } Regards,
} } Wharton
} }
} } *************************************************
} } Wharton Sinkler, Ph.D.
} } UOP LLC
} } 25 E. Algonquin Rd.
} } Des Plaines, IL 60017-5017
} } 847-391-3878
} }
} }
} } } -----Original Message-----
} } } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } } Sent: Tuesday, June 04, 2002 3:36 PM
} } } To: microscopy-at-sparc5.microscopy.com
} } } Subject: Re: TEM imaging plate technology
} } }
} } }
} ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} -----------------------------------------------------------------------.
} } }
} } }
} } } Dear Donald
} } }
} } } Before I bought the digital camera, I was looking on that image plate
} } } technology. It seems to me, it does not have serious advantage for
} most
} } } of
} } } us. I would like to concentrate my findings in a few sentences.
} } }
} } } Image plate PRO:
} } } - has more pixels that most digital cameras (some top-end cameras
} } } HAS similar amount).
} } } - has dynamic range same as digital cameras 12-16 bit.
} } } - it's more sensitive than film.
} } } - very good for electron diffraction experiments!!! (only one real
} plus to
} } } me).
} } } - manufacturers claimed that pixel resolution on the plate is
} comparable
} } } with film (it's difficult to proof because you have to scan film to
} } } compare, so it'll depend from the film scanner).
} } }
} } } Image plate CONTRA:
} } } Imitate the film procedure - you have to perform all procedures as for
} } } film, load/upload plates into cassettes (in the dark), change the
} magazine
} } }
} } } (wait for vacuum), load plates into the scanner (in the dark I
} believe),
} } } wait for scanning - 2 (or more, don't remember, up to 5 at full
} resolution
} } }
} } } I believe) min etc. So, it does not eliminate the dark-room, scanning
} is
} } } slow and then you have to process/save/print data. Time consuming.
} You
} } } have all disadvantages the classical film use: plate may be scratched
} } } during loading/uploading, deformed, lost, dropped on floor with
} valuable
} } } image etc. One plate is $100 I believe. No practical use in most
} } } biological applications I believe. I am not warranty that all my
} } } information is current, I was shopping for image plate system a few
} years
} } } ago and my comments represent the situation of that time. Sorry,
} } } manufacturing gays. Sergey
} } }
} } } At 08:07 AM 6/4/02 -0500, you wrote:
} } }
} } ------------------------------------------------------------------------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } } } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } } } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} } -----------------------------------------------------------------------.
} } } }
} } } }
} } } }
} } } }
} } } } Dear EM Netters,
} } } }
} } } } Does anyoun out there have experience with imaging plate technology
} (i.e.
} } } } Ditabis)? I
} } } } would like to hear pros and cons concidering this technology. How
} does
} } } it
} } } } compare to
} } } } digital CCD camera systems (cost, time, supplies, reliability. We
} are
} } } } looking into possibly
} } } } retrofitting our JEOL 100CX with an off-line system.
} } } }
} } } }
} } } } Thanks in advance
} } } }
} } } }
} } } } Donald G. Awbrey, HT (ASCP), QIHC
} } } } Electron Microscopy / Image Analysis
} } } } 817-878-5647
} } } } donaldawbrey-at-texashealth.org
} } } }
} } }
} } } ------------------------------------------------------
} } }
} } } Sergey Ryazantsev, Ph.D.
} } } Electron Microscopy
} } } Department of Biological Chemistry, School of Medicine
} } } University of California, Los Angeles
} } } Box 951737
} } } Los Angeles, CA 90095-1737
} } }
} } } (310) 825-1144 (office)
} } } Pager: (310) 845-0248
} } } FAX: (310) 206-5272 (departmental)
} } } mailto:sryazant-at-ucla.edu
} } }
}
} _____________________________________
}
} Sergey Ryazantsev Ph. D.
} Electron Microscopy
} UCLA School of Medicine
} Department of Biological Chemistry
} Box 951737
} Los Angeles, CA 90095-1737
}
} Phone: (310) 825-1144
} FAX (departmental): (310) 206-5272
} mailto:sryazant-at-ucla.edu
}
}
}


From daemon Wed Jun 5 14:01:35 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Wed, 05 Jun 2002 12:12:31 -0700
Subject: virus alert

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I just got message with subject "Subject: To call in the service
representative." from microscopy-at-sparc5.microscopy.com with attachment
aa_160x100[1].exe

This message contains some virus. Unfortunately my Norton Antivirus
eliminate it so quickly and I have no chance to find what virus it was.
Thanks for such nice gift. May be it's sort of sign, I am talking too much
on ListServer? Sergey
------------------------------------------------------

Sergey Ryazantsev, Ph.D.
Electron Microscopy
Department of Biological Chemistry, School of Medicine
University of California, Los Angeles
Box 951737
Los Angeles, CA 90095-1737

(310) 825-1144 (office)
Pager: (310) 845-0248
FAX: (310) 206-5272 (departmental)
mailto:sryazant-at-ucla.edu



From daemon Wed Jun 5 15:22:08 2002



From: Ingo Daberkow :      ingo.daberkow-at-tvips.com
Date: Wed, 05 Jun 2002 22:13:44 +0200
Subject: Re: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sergey,

Just another remark regarding the dynamic range. The imaging plates have a much
higher dynamic range than other electron detectors (up to 20 bit after several
scans). But please notice the dynamic range is defined as: (maximum signal/RMS
noise) – this is not the digitization level of the analog-to-digital converter.
This extreme high dynamic range is only necessary for the acquisition of
diffraction patterns. A “normal” bright-field image has typically a dynamic
range of about 100, which is mainly limited by the noise of the electrons
(shot-noise). This means, for such applications CCD cameras have the advantage
of a real-time acquisition (WYSIWYG) and furthermore the camera can be used for
checking or controlling of the TEM parameters (defocus, astigmatism, etc.)

Personally I have found the literature list of Wharton very useful – which
enables the interesting reader to collect more specific information.

Best wishes,
Ingo

+++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Dr. Ingo Daberkow
Tietz Video and Image Processing Systems GmbH
Eremitenweg 1
D-82131 Gauting, Germany
Tel: +49-89-8506567
FAX: +49-89-8509488
Internet: http://www.tvips.com/
Email: ingo.daberkow-at-tvips.com
+++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++

,



"Sinkler, Wharton" wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} OK, Truce Sergey!
}
} Your point is well taken, and yes, design considerations and convenience may
} overwhelm a good MTF or linearity.
}
} However, I think you are mistaken about dynamic range being similar for
} Imaging Plates and CCD's. That being said, I have gone back to look at the
} papers I mentioned and I'm not able to find a clear statement of what I
} mean.
}
} I believe (maybe someone will set us straight on this) that the saturation
} of the IP is considerably higher than the CCD. By 'saturation' I mean the
} dose (electrons/unit area) at which there is significant deviation from
} linearity.
}
} The reason saturation level is important (and why the IP is great for
} diffraction) is that signal/noise at low electron doses is dominated by shot
} noise. So if you are trying to record a diffraction pattern with a CCD with
} a single exposure without saturating the CCD at the strong spots, the weak
} reflections will be noisy simply because relatively few electrons are
} arriving there (statistical or 'shot' noise). With a higher saturation (I
} believe the case for IP), you can expose longer and get better statistics
} overall. (Of course you can also record multiple exposures with the CCD at
} different times and merge the data).
}
} Regards,
} Wharton
}
} } -----Original Message-----
} } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } Sent: Tuesday, June 04, 2002 7:42 PM
} } To: Microscopy-at-sparc5.microscopy.com
} } Subject: RE: TEM imaging plate technology
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Dear Wharton
} }
} } Are you seriously think, people will read that very specific papers? Do
} } you think, so many of us is ready to go through all this math etc? I
} } agree,
} } it's very important to understand the scientific foundation for device you
} }
} } are going to use. But, in practice you have deal not with such nice
} } things
} } as PSF (point spread function) or MTF. The camera with very nice PSF may
} } have terrible design and not suitable for use. Company, who manufactured
} } that baby may not respond on your telephone calls and finally you may
} } figure out that it does not exists anymore. I do read scientific
} } journals
} } and it's extremely easy to find informations on the Internet nova days,
} } ListServer to me is a place where we have chance to share our personal
} } experience (bad or good), which mostly is not published and not indexed by
} }
} } Medline/CC whatever. Anyway, thank you for the reference, I enjoyed
} } reading it. Sergey
} }
} } At 01:41 PM 6/4/02, you wrote:
} }
} } } Sergey et al,
} } }
} } } I'd just like to point out that instead of relying on uncertain
} } } recollections, there is some good work in the literature which discusses
} } } both slow-scan CCD's and imaging plates. Here's a start:
} } }
} } } J. M. Zuo "Electron detection characteristics of slow-scan CCD camera"
} } } Ultramicroscopy 66 (1996) 21-33.
} } }
} } } J. M. Zuo, M. R. McCartney and J. C. H. Spence "Performance of imaging
} } } plates for electron recording" Ultramicroscopy 66 (1996) 35-47
} } }
} } } G. Y. Fan and M. H. Ellisman "Digital imaging in transmission electron
} } } microscopy" Journal of Microscopy 200 (2000) 1-13.
} } }
} } } Although the technology is continuously evolving, the basic points are
} } still
} } } valid.
} } }
} } } Regards,
} } } Wharton
} } }
} } } *************************************************
} } } Wharton Sinkler, Ph.D.
} } } UOP LLC
} } } 25 E. Algonquin Rd.
} } } Des Plaines, IL 60017-5017
} } } 847-391-3878
} } }
} } }
} } } } -----Original Message-----
} } } } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } } } Sent: Tuesday, June 04, 2002 3:36 PM
} } } } To: microscopy-at-sparc5.microscopy.com
} } } } Subject: Re: TEM imaging plate technology
} } } }
} } } }
} } ------------------------------------------------------------------------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } America
} } } } To Subscribe/Unsubscribe -- Send Email to
} } ListServer-at-MSA.Microscopy.Com
} } } } On-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } }
} } -----------------------------------------------------------------------.
} } } }
} } } }
} } } } Dear Donald
} } } }
} } } } Before I bought the digital camera, I was looking on that image plate
} } } } technology. It seems to me, it does not have serious advantage for
} } most
} } } } of
} } } } us. I would like to concentrate my findings in a few sentences.
} } } }
} } } } Image plate PRO:
} } } } - has more pixels that most digital cameras (some top-end cameras
} } } } HAS similar amount).
} } } } - has dynamic range same as digital cameras 12-16 bit.
} } } } - it's more sensitive than film.
} } } } - very good for electron diffraction experiments!!! (only one real
} } plus to
} } } } me).
} } } } - manufacturers claimed that pixel resolution on the plate is
} } comparable
} } } } with film (it's difficult to proof because you have to scan film to
} } } } compare, so it'll depend from the film scanner).
} } } }
} } } } Image plate CONTRA:
} } } } Imitate the film procedure - you have to perform all procedures as for
} } } } film, load/upload plates into cassettes (in the dark), change the
} } magazine
} } } }
} } } } (wait for vacuum), load plates into the scanner (in the dark I
} } believe),
} } } } wait for scanning - 2 (or more, don't remember, up to 5 at full
} } resolution
} } } }
} } } } I believe) min etc. So, it does not eliminate the dark-room, scanning
} } is
} } } } slow and then you have to process/save/print data. Time consuming.
} } You
} } } } have all disadvantages the classical film use: plate may be scratched
} } } } during loading/uploading, deformed, lost, dropped on floor with
} } valuable
} } } } image etc. One plate is $100 I believe. No practical use in most
} } } } biological applications I believe. I am not warranty that all my
} } } } information is current, I was shopping for image plate system a few
} } years
} } } } ago and my comments represent the situation of that time. Sorry,
} } } } manufacturing gays. Sergey
} } } }
} } } } At 08:07 AM 6/4/02 -0500, you wrote:
} } } }
} } } ------------------------------------------------------------------------
} } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } America
} } } } } To Subscribe/Unsubscribe -- Send Email to
} } ListServer-at-MSA.Microscopy.Com
} } } } } On-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } }
} } } -----------------------------------------------------------------------.
} } } } }
} } } } }
} } } } }
} } } } }
} } } } } Dear EM Netters,
} } } } }
} } } } } Does anyoun out there have experience with imaging plate technology
} } (i.e.
} } } } } Ditabis)? I
} } } } } would like to hear pros and cons concidering this technology. How
} } does
} } } } it
} } } } } compare to
} } } } } digital CCD camera systems (cost, time, supplies, reliability. We
} } are
} } } } } looking into possibly
} } } } } retrofitting our JEOL 100CX with an off-line system.
} } } } }
} } } } }
} } } } } Thanks in advance
} } } } }
} } } } }
} } } } } Donald G. Awbrey, HT (ASCP), QIHC
} } } } } Electron Microscopy / Image Analysis
} } } } } 817-878-5647
} } } } } donaldawbrey-at-texashealth.org
} } } } }
} } } }
} } } } ------------------------------------------------------
} } } }
} } } } Sergey Ryazantsev, Ph.D.
} } } } Electron Microscopy
} } } } Department of Biological Chemistry, School of Medicine
} } } } University of California, Los Angeles
} } } } Box 951737
} } } } Los Angeles, CA 90095-1737
} } } }
} } } } (310) 825-1144 (office)
} } } } Pager: (310) 845-0248
} } } } FAX: (310) 206-5272 (departmental)
} } } } mailto:sryazant-at-ucla.edu
} } } }
} }
} } _____________________________________
} }
} } Sergey Ryazantsev Ph. D.
} } Electron Microscopy
} } UCLA School of Medicine
} } Department of Biological Chemistry
} } Box 951737
} } Los Angeles, CA 90095-1737
} }
} } Phone: (310) 825-1144
} } FAX (departmental): (310) 206-5272
} } mailto:sryazant-at-ucla.edu
} }
} }
} }



From daemon Wed Jun 5 15:38:10 2002



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Wed, 05 Jun 2002 15:31:51 -0500
Subject: Re: virus alert

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Did the message come with the standard MSA header? I doubt it, but it
should if it really came via the list. (Right, Nestor?) I have not received
the message you described.

I have had a number of virus reports sent back to me that a message of mine
had a virus that was caught by one of the growing number of corporate virus
checkers. My name was listed as the alleged sender of the message; however,
the return path listed a different account. I remember reading something on
the McAfee site that such is the behavior of several of the viruses out there.

So, some virus probably is spoofing the MSA address. Nestor has done a good
job of stopping most of the trash from getting through.

Warren

At 12:12 PM 6/5/02 -0700, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From daemon Wed Jun 5 16:00:45 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Wed, 05 Jun 2002 13:54:23 -0700
Subject: Re: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ingo
Thanks for comments. I absolutely agree with folks who mentioned extremely
wide dynamic range for image plates. I think there is some misunderstood
happening: In order to get data from image plate, you have to scan it and
record it as a digital image. The quality of scanners is different, some
of them could not provide 20 bit gray scale. When I pointed, that dynamic
range is the same for the images from CCD and IP I mean that the images
were recorded in 12-16 bit format, which is very usual for nova days. 20
bit format is non-readable by most software and therefore useless for such
simple person like me, who is still using Photoshop for most jobs. So,
from this point of view, we have practically the same dynamic range of the
final digital image. I am sorry if my posting confuse somebody. Image
plate technology widely used in X-ray crystallography and as a substitution
for X-ray film in molecular biology applications for decades. No question
about that: this is a great technology. BUT, my point, is that current
realization of this technology in EM does not have serious advantages over
normal film/digital camera combination. AND: keep in mind, how long you
will scan to get those 20 bits dynamic range... Thanks again for the very
good point. Sergey


At 01:13 PM 6/5/02, you wrote:
} Dear Sergey,
}
} Just another remark regarding the dynamic range. The imaging plates have a
} much
} higher dynamic range than other electron detectors (up to 20 bit after several
} scans). But please notice the dynamic range is defined as: (maximum signal/RMS
} noise) ­ this is not the digitization level of the analog-to-digital
} converter.
} This extreme high dynamic range is only necessary for the acquisition of
} diffraction patterns. A “normal” bright-field image has typically a dynamic
} range of about 100, which is mainly limited by the noise of the electrons
} (shot-noise). This means, for such applications CCD cameras have the advantage
} of a real-time acquisition (WYSIWYG) and furthermore the camera can be
} used for
} checking or controlling of the TEM parameters (defocus, astigmatism, etc.)
}
} Personally I have found the literature list of Wharton very useful ­ which
} enables the interesting reader to collect more specific information.
}
} Best wishes,
} Ingo
}
} +++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
} Dr. Ingo Daberkow
} Tietz Video and Image Processing Systems GmbH
} Eremitenweg 1
} D-82131 Gauting, Germany
} Tel: +49-89-8506567
} FAX: +49-89-8509488
} Internet: http://www.tvips.com/
} Email: ingo.daberkow-at-tvips.com
} +++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
}
} ,
}
}
}
} "Sinkler, Wharton" wrote:
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} } OK, Truce Sergey!
} }
} } Your point is well taken, and yes, design considerations and
} convenience may
} } overwhelm a good MTF or linearity.
} }
} } However, I think you are mistaken about dynamic range being similar for
} } Imaging Plates and CCD's. That being said, I have gone back to look at the
} } papers I mentioned and I'm not able to find a clear statement of what I
} } mean.
} }
} } I believe (maybe someone will set us straight on this) that the saturation
} } of the IP is considerably higher than the CCD. By 'saturation' I mean the
} } dose (electrons/unit area) at which there is significant deviation from
} } linearity.
} }
} } The reason saturation level is important (and why the IP is great for
} } diffraction) is that signal/noise at low electron doses is dominated by
} shot
} } noise. So if you are trying to record a diffraction pattern with a CCD
} with
} } a single exposure without saturating the CCD at the strong spots, the weak
} } reflections will be noisy simply because relatively few electrons are
} } arriving there (statistical or 'shot' noise). With a higher saturation (I
} } believe the case for IP), you can expose longer and get better statistics
} } overall. (Of course you can also record multiple exposures with the CCD at
} } different times and merge the data).
} }
} } Regards,
} } Wharton
} }
} } } -----Original Message-----
} } } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } } Sent: Tuesday, June 04, 2002 7:42 PM
} } } To: Microscopy-at-sparc5.microscopy.com
} } } Subject: RE: TEM imaging plate technology
} } }
} } } ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------.
} } }
} } }
} } } Dear Wharton
} } }
} } } Are you seriously think, people will read that very specific papers? Do
} } } you think, so many of us is ready to go through all this math etc? I
} } } agree,
} } } it's very important to understand the scientific foundation for
} device you
} } }
} } } are going to use. But, in practice you have deal not with such nice
} } } things
} } } as PSF (point spread function) or MTF. The camera with very nice PSF may
} } } have terrible design and not suitable for use. Company, who manufactured
} } } that baby may not respond on your telephone calls and finally you may
} } } figure out that it does not exists anymore. I do read scientific
} } } journals
} } } and it's extremely easy to find informations on the Internet nova days,
} } } ListServer to me is a place where we have chance to share our personal
} } } experience (bad or good), which mostly is not published and not
} indexed by
} } }
} } } Medline/CC whatever. Anyway, thank you for the reference, I enjoyed
} } } reading it. Sergey
} } }
} } } At 01:41 PM 6/4/02, you wrote:
} } }
} } } } Sergey et al,
} } } }
} } } } I'd just like to point out that instead of relying on uncertain
} } } } recollections, there is some good work in the literature which discusses
} } } } both slow-scan CCD's and imaging plates. Here's a start:
} } } }
} } } } J. M. Zuo "Electron detection characteristics of slow-scan CCD camera"
} } } } Ultramicroscopy 66 (1996) 21-33.
} } } }
} } } } J. M. Zuo, M. R. McCartney and J. C. H. Spence "Performance of imaging
} } } } plates for electron recording" Ultramicroscopy 66 (1996) 35-47
} } } }
} } } } G. Y. Fan and M. H. Ellisman "Digital imaging in transmission electron
} } } } microscopy" Journal of Microscopy 200 (2000) 1-13.
} } } }
} } } } Although the technology is continuously evolving, the basic points are
} } } still
} } } } valid.
} } } }
} } } } Regards,
} } } } Wharton
} } } }
} } } } *************************************************
} } } } Wharton Sinkler, Ph.D.
} } } } UOP LLC
} } } } 25 E. Algonquin Rd.
} } } } Des Plaines, IL 60017-5017
} } } } 847-391-3878
} } } }
} } } }
} } } } } -----Original Message-----
} } } } } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } } } } Sent: Tuesday, June 04, 2002 3:36 PM
} } } } } To: microscopy-at-sparc5.microscopy.com
} } } } } Subject: Re: TEM imaging plate technology
} } } } }
} } } } }
} } } ------------------------------------------------------------------------
} } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } } America
} } } } } To Subscribe/Unsubscribe -- Send Email to
} } } ListServer-at-MSA.Microscopy.Com
} } } } } On-Line Help
} } } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } }
} } } -----------------------------------------------------------------------.
} } } } }
} } } } }
} } } } } Dear Donald
} } } } }
} } } } } Before I bought the digital camera, I was looking on that image plate
} } } } } technology. It seems to me, it does not have serious advantage for
} } } most
} } } } } of
} } } } } us. I would like to concentrate my findings in a few sentences.
} } } } }
} } } } } Image plate PRO:
} } } } } - has more pixels that most digital cameras (some top-end cameras
} } } } } HAS similar amount).
} } } } } - has dynamic range same as digital cameras 12-16 bit.
} } } } } - it's more sensitive than film.
} } } } } - very good for electron diffraction experiments!!! (only one real
} } } plus to
} } } } } me).
} } } } } - manufacturers claimed that pixel resolution on the plate is
} } } comparable
} } } } } with film (it's difficult to proof because you have to scan film to
} } } } } compare, so it'll depend from the film scanner).
} } } } }
} } } } } Image plate CONTRA:
} } } } } Imitate the film procedure - you have to perform all procedures
} as for
} } } } } film, load/upload plates into cassettes (in the dark), change the
} } } magazine
} } } } }
} } } } } (wait for vacuum), load plates into the scanner (in the dark I
} } } believe),
} } } } } wait for scanning - 2 (or more, don't remember, up to 5 at full
} } } resolution
} } } } }
} } } } } I believe) min etc. So, it does not eliminate the dark-room,
} scanning
} } } is
} } } } } slow and then you have to process/save/print data. Time consuming.
} } } You
} } } } } have all disadvantages the classical film use: plate may be scratched
} } } } } during loading/uploading, deformed, lost, dropped on floor with
} } } valuable
} } } } } image etc. One plate is $100 I believe. No practical use in most
} } } } } biological applications I believe. I am not warranty that all my
} } } } } information is current, I was shopping for image plate system a few
} } } years
} } } } } ago and my comments represent the situation of that time. Sorry,
} } } } } manufacturing gays. Sergey
} } } } }
} } } } } At 08:07 AM 6/4/02 -0500, you wrote:
} } } } }
} } } } ------------------------------------------------------------------------
} } } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } } America
} } } } } } To Subscribe/Unsubscribe -- Send Email to
} } } ListServer-at-MSA.Microscopy.Com
} } } } } } On-Line Help
} } } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } }
} } } } -----------------------------------------------------------------------.
} } } } } }
} } } } } }
} } } } } }
} } } } } }
} } } } } } Dear EM Netters,
} } } } } }
} } } } } } Does anyoun out there have experience with imaging plate technology
} } } (i.e.
} } } } } } Ditabis)? I
} } } } } } would like to hear pros and cons concidering this technology. How
} } } does
} } } } } it
} } } } } } compare to
} } } } } } digital CCD camera systems (cost, time, supplies, reliability. We
} } } are
} } } } } } looking into possibly
} } } } } } retrofitting our JEOL 100CX with an off-line system.
} } } } } }
} } } } } }
} } } } } } Thanks in advance
} } } } } }
} } } } } }
} } } } } } Donald G. Awbrey, HT (ASCP), QIHC
} } } } } } Electron Microscopy / Image Analysis
} } } } } } 817-878-5647
} } } } } } donaldawbrey-at-texashealth.org
} } } } } }
} } } } }
} } } } } ------------------------------------------------------
} } } } }
} } } } } Sergey Ryazantsev, Ph.D.
} } } } } Electron Microscopy
} } } } } Department of Biological Chemistry, School of Medicine
} } } } } University of California, Los Angeles
} } } } } Box 951737
} } } } } Los Angeles, CA 90095-1737
} } } } }
} } } } } (310) 825-1144 (office)
} } } } } Pager: (310) 845-0248
} } } } } FAX: (310) 206-5272 (departmental)
} } } } } mailto:sryazant-at-ucla.edu
} } } } }
} } }
} } } _____________________________________
} } }
} } } Sergey Ryazantsev Ph. D.
} } } Electron Microscopy
} } } UCLA School of Medicine
} } } Department of Biological Chemistry
} } } Box 951737
} } } Los Angeles, CA 90095-1737
} } }
} } } Phone: (310) 825-1144
} } } FAX (departmental): (310) 206-5272
} } } mailto:sryazant-at-ucla.edu
} } }
} } }
} } }

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Wed Jun 5 16:03:48 2002



From: Larry Ackerman :      mishot-at-itsa.ucsf.edu
Date: Wed, 05 Jun 2002 14:02:42 -0700
Subject: CCD FOR FRET

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We hope to purchase a new sensitive CCD and one future application is FRET.
I have no experience with this technique. Please advise me about the time
between the two acquired images, ie. how fast should the system operate?
Thanks

Larry Ackerman
Howard Hughes Medical Institute
UCSF, Box 0725, Rm U226
533 Parnassus Avenue
San Francisco, CA 94143 USA

415-476-8751 FAX 415-476-5774

http://www.ucsf.edu/jan



From daemon Wed Jun 5 16:15:42 2002



From: Hong Yi :      hyi-at-emory.edu
Date: Wed, 5 Jun 2002 17:09:38 -0400 (EDT)
Subject: Embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All:
I recently did a pre-embedding double immunolabeling run on 50 um
rat brain vibratome sections. The labeling looked beautiful on LM so I
flat embedded sections for EM. Our standard protocol for resin
infiltration and embedding involves placing sections in a mixture of
propylene oxide and Eponate 12 (Pella) at 1:1 ratio in a aluminum dish
overnight, followed by placing sections in 100% Epon for a few hours, and
then embedding with fresh Epon. It has always worked fine for us.
However, I used Epon resin from a different source this time, and am
having problem sectioning. Thin sections disintegrate once they are
floating on water. I have tried to leave embedded tissue in oven for a
few extra days and radiate blocks with UV light. But so far nothing
helped. Does anyone have any suggestion on what else I can try to rescue
these embedded sections, or I should just give it up and start another
run?


Hong



From daemon Wed Jun 5 17:08:52 2002



From: Scott Henderson :      Scott.Henderson-at-mssm.edu
Date: Wed, 05 Jun 2002 18:01:19 -0400
Subject: technical position available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



A technical position is available in the Microscopy Shared Resource
Facility at Mount Sinai School of Medicine in New York. The
successful candidate will participate in microscopy studies of
various biological systems. Duties include instructing and assisting
users of the facility, sample preparation, image analysis and minor
equipment maintenance. Applicants should have excellent
communication and organizational skills, an understanding of basic
laboratory procedures, and the ability to manage a large and varied
workload. Qualifications include a degree in Biology/Life Sciences,
experience with laser scanning microscopy (confocal and/or
multi-photon), sample preparation (e.g. immunofluorescence), digital
imaging, image analysis and routine equipment maintenance. Computer
skills are essential. Some experience with electron microscopy is an
asset.

For consideration, please mail/email your resume to:

Scott Henderson, Ph.D.,
Director, Microscopy Shared Resource Facility,
Box 1007,
Mount Sinai School of Medicine,
One Gustave L. Levy Place,
New York, NY 10029-6574

email: Scott.Henderson-at-mssm.edu

"We are an equal opportunity employer and foster diversity in the workplace."



From daemon Wed Jun 5 17:32:05 2002



From: Bruce Girrell :      bigirrell-at-microlinetc.com
Date: Wed, 5 Jun 2002 18:28:37 -0400
Subject: RE: repair for old ISI sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Steve,

While looking at the operation manual for the E5000, I saw that the machine
will not start operation until the pirani gauge goes below 0.2 Torr.
Assuming that you pumped your system down to at least that level, then
something is wrong with the pressure relay that controls the high voltage or
some part of that circuit is not tripping the relay.

The manual says that you should be able to hear a click when the relay kicks
in at about 0.2 Torr.

Bruce Girrell



From daemon Wed Jun 5 18:39:42 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Wed, 5 Jun 2002 17:38:41 -0600
Subject: virus alert

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Relax, Sergey,

you were probably hit by the W32.Klez.H-at-mm worm or virus. As a twist it does
not use the real senders email address, but "fakes" the email address of
some other person in the senders email list. So you got the email from
someone, who has the listserver somewhere in his address book or wherever
the worm gets its information. In other words: This email came from John
Doe, who at some point had contact with the listserver and also has your
email address somewhere.

I don't think you can get the worm through the list server, to my knowledge
it requires some form of html email, which is not accepted on the server. Is
that right, Nestor?

mike

} } } } } } } } } } WE HAVE MOVED { { { { { { { { {
please make a note of the new address below

Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Sergey Ryazantsev [mailto:sryazant-at-ucla.edu]
Sent: Wednesday, June 05, 2002 1:13 PM
To: microscopy-at-sparc5.microscopy.com


I just got message with subject "Subject: To call in the service
representative." from microscopy-at-sparc5.microscopy.com with attachment
aa_160x100[1].exe

This message contains some virus. Unfortunately my Norton Antivirus
eliminate it so quickly and I have no chance to find what virus it was.
Thanks for such nice gift. May be it's sort of sign, I am talking too much
on ListServer? Sergey
------------------------------------------------------

Sergey Ryazantsev, Ph.D.
Electron Microscopy
Department of Biological Chemistry, School of Medicine
University of California, Los Angeles
Box 951737
Los Angeles, CA 90095-1737

(310) 825-1144 (office)
Pager: (310) 845-0248
FAX: (310) 206-5272 (departmental)
mailto:sryazant-at-ucla.edu



From daemon Wed Jun 5 19:52:01 2002



From: Gary Gaugler :      gary-at-gaugler.com
Date: Wed, 05 Jun 2002 17:46:40 -0700
Subject: Re: virus alert

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Set up Norton to put any virus into quarantine at your
request. Each msg that has a virus will come up with
a Norton warning screen and ask if you want to quarantine
the msg (assuming that repair failed). If the virus could
be repaired, this msg stream does not appear. That could
explain why things happened so quickly.

I am using NAV System Works 5.02 (2002) and it works
great. Be sure to keep the virus definitions updated via
live update (tm).

gary g.


At 12:12 PM 6/5/2002, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Wed Jun 5 20:44:30 2002



From: Helvey, Marc :      Marc.Helvey-at-vlsistd.com
Date: Wed, 5 Jun 2002 18:37:49 -0700
Subject: AFM calibration - standards and procedures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Matthew -

http://www.vlsistandards.com/products/so_surfTopStan.asp?cid=3&sid=47

Please see the traceability chart at the bottom of the page. Feel free to
contact me offline about this product.

Regards -

Marc Helvey
Strategic Accounts Manager
VLSI Standards, Inc.
3087 North First Street
San Jose, CA 95134-2006
Phone: (408) 428-1800, ext. 108
FAX: (408) 428-9555
Mobile: (408) 221-7581
E-Mail: marc.helvey-at-vlsistd.com {mailto:marc.helvey-at-vlsistd.com}
Internet: http://www.vlsistandards.com




-----Original Message-----
} From: Stephenson, Matthew [mailto:stephenson-at-impactanalytical.com]
Sent: Wednesday, June 05, 2002 9:40 AM
To: Microscopy-at-sparc5.microscopy.com


Greetings all!

I'm in the process of resuscitating a TopoMetrix TMX 2000 with Discoverer
stage for contact and non-contact AFM. We use three scanners: a 70 micron
tripod, a 7 micron tube, and a 0.76 micron tube. The instrument is up and
running, so the time has come to calibrate, incorporate into our quality
program, and get some customers. My questions are as follows:

1. What reference standards are recommended for x, y, and z distance
calibration? Is anything available that is NIST traceable, or barring that,
certifiable by a reputable authority?

2. Have widely accepted AFM calibration procedures emerged in the
microscopy community? I'd be particularly interested in citations for sound
calibration procedure, and hoary wisdom regarding the effect of instrument
variables on the calibration process and accuracy.

I look forward to your comments!

Matthew K. Stephenson
Analytical Associate
Impact Analytical
1910 West Saint Andrews Road
Midland, MI 48640
(989) 832-5555 X506
stephenson-at-impactanalytical.com




From daemon Wed Jun 5 21:01:28 2002



From: georgiana_sprott-at-aemail4u.com
Date: Wed, 05 Jun 0102 23:27:40 +0200
Subject: BIZ, .INFO, .COM for only $14.95

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
charset="iso-8859-1"
Content-Transfer-Encoding: 8bit
X-Priority: 3 (Normal)
X-MSMail-Priority: Normal
X-Mailer: The Bat! (v1.52f) Business
Importance: Normal



From: georgiana_sprott-at-aemail4u.com
Date: Wed, 05 Jun 0102 23:27:40 +0200
Subject: BIZ, .INFO, .COM for only $14.95

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America



LATEST NEWS:

The new domain names are finally available to the general public at discount prices. Now you can register one of the exciting new .BIZ or .INFO domain names, as well as the original .COM and .NET names for just $14.95. These brand new domain extensions were recently approved by ICANN and have the same rights as the original .COM and .NET domain names. The biggest benefit is of-course that the .BIZ and .INFO domain names are currently more available. i.e. it will be much easier to register an attractive and easy-to-remember domain name for the same price. Visit: http://www.affordable-domains.com today for more info.

Register your domain name today for just $14.95 at: http://www.affordable-domains.com. Registration fees include full access to an easy-to-use control panel to manage your domain name in the future.

Sincerely,

Domain Administrator
Affordable Domains


To remove your email address from further promotional mailings from this company, click here:
http://www.centralremovalservice.com/cgi-bin/domain-remove.cgi
35

8647XXKc8-630lbAI4811QDZT1-070GUl30


From daemon Wed Jun 5 21:18:52 2002



From: Bill Miller :      microbill-at-mohawk.net
Date: Wed, 05 Jun 2002 22:10:01 -0400
Subject: RE: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id VAA17799
for dist-Microscopy; Wed, 5 Jun 2002 21:17:49 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id VAA17792
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Wed, 5 Jun 2002 21:17:19 -0500 (CDT)
Received: from mohegan.mohawk.net (mohegan.mohawk.net [63.66.68.21])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id VAA17784
for {Microscopy-at-sparc5.microscopy.com} ; Wed, 5 Jun 2002 21:17:07 -0500 (CDT)
Received: from MICROBILL.mohawk.net (ool-182dcdde.dyn.optonline.net [24.45.205.222])
(authenticated)
by mohegan.mohawk.net (8.11.4/8.11.3) with ESMTP id g562CNW87821;
Wed, 5 Jun 2002 22:12:23 -0400 (EDT)
Message-Id: {5.1.0.14.2.20020605204718.044b1c48-at-mail.mohawk.net}
X-Sender: microbill-at-mail.mohawk.net
X-Mailer: QUALCOMM Windows Eudora Version 5.1


First - ElectroImage will probably be selling the Ditabis Imaging Plate
system - but even if that were not true there should be some reality check
in all this - It is beginning to sounds very suspiciously like "Oh my god ,
we can't have another competitor." There are times when the plusses of a
CCD system are the right answer for the TEM digitization problem. Other
times there may actually be some different solution - scan the negatives,
shoot small portions of the negatives with a digital camera and a light
box, or maybe even use an imaging plate system. "One size does not fit
all." If it's all you have to sell, then it is right and proper to defend
it but to spread what you "think" without knowing is particularly bad
science and not particularly helpful to anyone. From the hail storm of
resistance though , you are not alone.


} Careful: the following is an opinion of a CCD system manufacturer. Reader
} discretion is advised!
}
} If I remember correctly, the manufacturers for he imaging plates claim a
} dynamic range of 20 bit, which is considerably more than the 12 or 14 bit
} from a CCD camera. If that, however, makes a big difference for diffraction
} patterns I don't know.

Then why bring it up?

} If you want to record the weakest spots, you need to
} expose at least for a time so that the signal is above noise. Since CCD
} cameras are very sensitive (down to single electron sensitivity), there is
} not much that can be improved by the imaging plates.

Imaging plates lack the readout noise of CCD's . It is true that there are
CCD's that can detect single photons but to my knowledge no one uses them
in this application. Certainly the new photon to electron cameras fit into
this category. Having higher bit depth means not only that you can "see"
the dimmer spots but that you can miss the exact exposure and still get
good images. The high DQE at low dose minimizes the exposure as does
the the large image area. Additionally, software like Lucis is ideally
suited to extract tremendous detail from just such high bit depth images.


} There are no half
} electrons (and if you see 1/3 electrons, book a flight to Stockholm).

It's DQE that counts here. Imaging plates have a DQE of .8 as low as
.001e/cm squared. What is the DQE of a CCD camera at low dose exposures?
The answer is



} If the
} 20 bits are sufficient to image the weakest beams together with the
} transmitted beam without saturation I don't know. With the Imaging plates
} you can probably get the weakest spots and more medium intensity spots at
} the same time than with a CCD. With a CCD you have the advantage that you
} immediately see the result and can adjust your exposure accordingly.


And if the sample is beam sensitive, then what? Again, there is no one
right answer.

} It's
} also very easy to take exposures at different exposure times and combine
} them, as the CCDs are very linear.


The imaging plates are linear over their six order of magnitude range -
CCDs as well as film has a sigmoid curve response and are linear only over
the central portion of their total response range.


} I can't remember anything about linearity of the imaging plates. But the way
} they work (excite crystals into a semi-stable state through the light from a
} phosphor and then "read' them by relaxing into ground stage through a
} laser), I would think there are some non-linearities as with increasing
} exposure you have less and less crystallites that can be excited. Whether
} that's critical for quantitative measurements I can't say.


And, indeed, you are quite wrong.


} Finally, the images on the plates are stable for a few hours. So you can't
} leave the plates in for a few days if you want the best results.

That is correct.



} One area, where in my experience (as a TEM user) CCD cameras have a big
} advantage is dark field imaging. I remember taking several films with
} different exposures to make sure that I had one that was right. With a CCD
} there is no guesswork anymore, as you can see the images immediately.


The extended dynamic range of the imaging plates pretty much negates this
re-exposure scheme.


} Resolution: theoretically the plate film should have a worse resolution that
} film. The electron beam has to travel through the crystallite layer, create
} light in the phosphor layer, and the photons then have to travel back to the
} crystallites to excite them. This should add to the dispersion of the beam,
} resulting in a reduced resolution. The manufacturers of the imaging plates
} claim around 20 microns resolution. If that is a 20 micron resolution at 20
} bit dynamic range they don't say.


15um resolution with .4MTF at the Nyquest frequency


} Again, we produce those CCD systems, and I definitely have a bias. But I
} thought I put my 2 cents in.
}
} Mike


The imaging plates offer high resolution over a large area - not quite the
resolution of film but the same area and with better sensitivity, higher
dynamic range and much better linearity. With up to 6000 x 5000 pixels (
actually from 1800 x 1250 to 6000 x 5000) the imaging plates offer much
more resolution for much less money than any other commercially available
system. They are not perfect but they can represent a solution to replace
film to produce high quality digital images. My position is to try to offer
the best solution for the customers problem. There are times when a CCD
system is the absolute right solution, as pointed out by Dr. Daberkow, but
there are times when it may not be. Now there is a choice.

Bill Miller
ElectroImage

} } } } } } } } } } } WE HAVE MOVED { { { { { { { { {
} please make a note of the new address below
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 12596 West Bayaud Avenue
} Suite 300
} Lakewood, CO 80228
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} } From: Sinkler, Wharton [mailto:WSinkler-at-uop.com]
} Sent: Wednesday, June 05, 2002 7:23 AM
} To: 'Sergey Ryazantsev'; Microscopy-at-sparc5.microscopy.com
} Subject: RE: TEM imaging plate technology
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Wed Jun 5 21:39:48 2002



From: zaluzec-at-microscopy.com
Date: Wed, 5 Jun 2002 21:30:12 -0500
Subject: Administrivia: Virus Alert

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues

Yes, Mike Bode and the othere that have replied are correct.
The listserver is setup to reject any messages with attachments.
These include EXE, HTML, and a host of others all of which
carry virus or are suspect/junk mail. If any valid mail
is caught by the filter and it does happen. Please read the directions in the
"returned message" , follow the instructions and we
can get your message through to the list.

The latest virus W32.Klez.H-at-mm worm, and variants there of spoof (fake)
email addresses and make someone else look as if someone else is
the "bad guy". This is likely the case with Sergey's mail.

As Warren has pointed out all Listserver mail will
have the banner/heading in the body of the message.
If you get suspect mail which carries the listserver
address and DO NOT see that banner it is guarenteed
to be a spoofed message. However, that being said,
the filters are not perfect and some spam gets through,
like the one on WWW Sites today, which I've now
added to my ever growing spam site list .

Just in case your curious the spam filter stopped 102
junk mail messages in the last week from reaching
all of you... I'll warn you this is a ever growing problem
and I have to continuously fight back the junk... It's
not what I would call fun..

Nestor
Your Friendly (and sometime tired) Neighborhood SysOp



From daemon Wed Jun 5 21:55:58 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Wed, 5 Jun 2002 20:56:46 -0600
Subject: RE: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bill,

I don't want to drag this out, you have your opinion and I have mine. But
let me correct you on one thing: CCDs are very linear over their range. If
you don't believe me, then perhaps you want to check out this site:

http://www.roper.co.jp/Html/teflin.htm

Here is a quote: High-performance CCD (HCCD) imagers have extremely good
linearity. Deviations from linearity are often less than a few tenths of a
percent for over five orders of magnitude.

Our measurements show similar results.

mike

} } } } } } } } } } WE HAVE MOVED { { { { { { { { {
please make a note of the new address below

Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Bill Miller [mailto:microbill-at-mohawk.net]
Sent: Wednesday, June 05, 2002 8:10 PM
To: Mike Bode; 'Microscopy-at-MSA.Microscopy.Com'


From daemon Wed Jun 5 22:41:19 2002



From: Gary Gaugler :      gary-at-gaugler.com
Date: Wed, 05 Jun 2002 20:36:15 -0700
Subject: Re: virus alert

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Attachments are not allowed on the MSA listserver. Nestor
does a great job of keeping postings true to this rule. If you
do have an attachment post that slips through, you probably
should update Nestor about it. Send your header info to him
if possible to help him differentiate your situation from others.

Nevertheless, Norton AV will trap any known virus in the body
or attachment of an e-mail message--or web-based thread.
Lately, I'm getting klez and sircam viruses. these are old and
well-known pathogens. Now, simply a pain.

Some simple (not free) actions for a virus-free environment is
to dump Microsoft Outlook and web browsers in favor of Qualcomm's
Eudora e-mail client and get Norton Anti-Virus (part of Norton System Works,
which includes other good options) and do not open attachments from
anyone you do not know. Beware of .SCR attachments (screen
savers). The central theme is to use Eudora for e-mail rather
than Netscape or IE5 or IE6 browsers or Outlook. You can also get relief
from viruses on the usenet by not using the browsers and instead,
using ForteInc Agent newsreader....very nice (using it for 5 years).

gary g.


At 12:12 PM 6/5/2002, you wrote:

} I just got message with subject "Subject: To call in the service
} representative." from microscopy-at-sparc5.microscopy.com with attachment
} aa_160x100[1].exe
}
} This message contains some virus. Unfortunately my Norton Antivirus
} eliminate it so quickly and I have no chance to find what virus it was.
} Thanks for such nice gift. May be it's sort of sign, I am talking too
} much on ListServer? Sergey
} ------------------------------------------------------
}
} Sergey Ryazantsev, Ph.D.
} Electron Microscopy
} Department of Biological Chemistry, School of Medicine
} University of California, Los Angeles
} Box 951737
} Los Angeles, CA 90095-1737
}
} (310) 825-1144 (office)
} Pager: (310) 845-0248
} FAX: (310) 206-5272 (departmental)
} mailto:sryazant-at-ucla.edu
}



From daemon Thu Jun 6 04:12:33 2002



From: Khalid Boulahya :      khalid-at-quim.ucm.es
Date: Thu, 06 Jun 2002 11:02:42 +0200
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe

we are looking urgently for a stage double tilt holder for EM400 family or
CM Philips microscopes (not compustage holder) for reasonable price.

best regards



From daemon Thu Jun 6 06:54:12 2002



From: Khalid Boulahya :      khalid-at-quim.ucm.es (by way of MicroscopyListserver)
Date: Thu, 6 Jun 2002 06:42:55 -0500
Subject: TEM Stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear friends,

we are looking urgently for a stage double tilt holder for EM400
family or CM Philips microscopes (not compustage holder) for
reasonable price.

best regards

**************************************************************
Khalid Boulahya
Depto de Química Inorgánica
Facultad De Químicas
Universidad Complutense de Madrid (UCM)
28040 Spain
khalid-at-quim.ucm.es
**************************************************************


From daemon Thu Jun 6 07:42:35 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Thu, 6 Jun 2002 08:35:21 -0400
Subject: RE: Embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Morning Hong,
The last time I had such a problem, I remember that I immediately
tried to section one of the blank blocks I always form from new batches of
resin. In my case, that block belied my feeling, after my first blank run
with the new resin, that I could trust the formulation of the new stuff. In
that instance, I had to go back and set up another run, after I practiced a
couple more times with the new furmulations.

My current rules: 1) Always make certain that a new type of
resin works by itself before you commit tissue to it. I don'/t like
sectioning blanks either, but the procedure helps to prevent REAL problems.

2) Always keep 1-5 10ml syringes of
pre-mixed resin from an old batch that really works so that when someone
comes along with a specimen that can't be replaced, you have a reliable
embedment to use.
3) Always mix your own stuff. My
experience has been that whenever a technician begins to mix things that
always work, you have to start looking for a replacement.

Sorry for your loss,

Fred Monson

Frederick C. Monson, PhD
Center for Advanced Scientific Imaging
Schmucker II Science Center
West Chester University
South Church Street and Rosedale
West Chester, Pennsylvania, USA, 19383
Phone: 610-738-0437
FAX: 610-738-0437
fmonson-at-wcupa.edu
CASI URL: http://darwin.wcupa.edu/casi/
WCUPA URL: http://www.wcupa.edu/
Visitors URL: http://www.wcupa.edu/_visitors/


} ----------
} From: Hong Yi
} Sent: Wednesday, June 5, 2002 5:09 PM
} To: microscopy-at-sparc5.microscopy.com
} Subject: Embedding
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear All:
} I recently did a pre-embedding double immunolabeling run on 50 um
} rat brain vibratome sections. The labeling looked beautiful on LM so I
} flat embedded sections for EM. Our standard protocol for resin
} infiltration and embedding involves placing sections in a mixture of
} propylene oxide and Eponate 12 (Pella) at 1:1 ratio in a aluminum dish
} overnight, followed by placing sections in 100% Epon for a few hours, and
} then embedding with fresh Epon. It has always worked fine for us.
} However, I used Epon resin from a different source this time, and am
} having problem sectioning. Thin sections disintegrate once they are
} floating on water. I have tried to leave embedded tissue in oven for a
} few extra days and radiate blocks with UV light. But so far nothing
} helped. Does anyone have any suggestion on what else I can try to rescue
} these embedded sections, or I should just give it up and start another
} run?
}
}
} Hong
}
}
}


From daemon Thu Jun 6 08:46:57 2002



From: Young, Gene (GP) :      GPYOUNG-at-dow.com
Date: Thu, 6 Jun 2002 09:38:51 -0400
Subject: Re:TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sergey,

Can you give more details on how your Gatan system is setup to broadcast
live over a network? Are you using something like Netmeeting? I am very
interested in doing this.

Gene

Gene P. Young
Research Technologist
Analytical Sciences, Polymer Characterization

The Dow Chemical Company
2301 N. Brazosport Blvd., B-1470
Freeport, Texas 77541-3257
Fax: (979) 238-0095
Phone: (979) 238-1579


-----Original Message-----
} From: Sergey Ryazantsev [mailto:sryazant-at-ucla.edu]
Sent: Monday, June 03, 2002 11:02 PM
To: Microscopy-at-sparc5.microscopy.com

{snip}
-another thing is academic: The live image from digital camera I broadcast
to the Departmental network, so you could see live image on any
departmental computer. PIs now comfortably sit in their offices and enjoy
watching how their students working on the TEM. Similarly it works for
Internet. We did a few session when our collaborators observe their
samples sitting far-far away from UCLA.

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Thu Jun 6 08:47:03 2002



From: Jamila Shawon :      Jamila_Shawon-at-student.uml.edu
Date: Thu, 06 Jun 2002 09:38:14 -0400
Subject: TEM sample prep

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all:

recently I am trying to prepare TEM sample of CMP polyurethane pads. I
made the sample with embaded resin and did microtoming. After
microtoming, I did staining by RuO4 (as my purpose is to identfy the
hard and soft phases). But my problem is that I have no idea how the
stained polyurethane will look like. I tried several times, but it was
hard to distinguish the sample from resin. So does anyone have any idea
about the PU cmp pads TEM images after staining? Please let me know.

Thanks
Jamila Shawon



From daemon Thu Jun 6 08:58:41 2002



From: Lou bustillos :      lbustillos-at-amalab.com
Date: Thu, 6 Jun 2002 09:52:35 -0400
Subject: Chillers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

First I would like to say thank you to everyone that helped me with my last
email about purchasing a carbon coater. Hopefully installation will be in
the next two weeks.

My new question is about chillers for a JEOL 100CX II. The state that we
are located in is beginning to have a drought and we want to do our part by
changing our water chillers to air cooled chillers. I would like to know
your experience with switching to an air cooled system. Also, what
manufacture would you recommend? I have two 100CX II and they will have
there own separate units. Thank you for your help.



From daemon Thu Jun 6 09:08:51 2002



From: Jim Romanow :      bsgphy3-at-uconnvm.uconn.edu
Date: Thu, 6 Jun 2002 10:03:44 -0400
Subject: TEM Stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Contact:

Optechs, Inc.
Used Philips parts
Dick Turnage
446 Acapesket Road
East Falmouth, MA 02536
617 739-7900 voice
978 667-1827 fax


}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

James S. Romanow
The University of Connecticut
Physiology and Neurobiology Department
Electron Microscopy Facility
Unit 2242
Storrs, CT 06269-2242
bsgphy3-at-uconnvm.uconn.edu
860 486-2914 voice
860 486-1936 fax




From daemon Thu Jun 6 10:03:19 2002



From: Ingo Daberkow :      ingo.daberkow-at-tvips.com
Date: Thu, 06 Jun 2002 16:54:33 +0200
Subject: Re: TEM imaging plate technology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sergey, Bill, Mike,

After I have been contacted by a listener off-line, I want to correct my remark:
“A normal bright-field image has typically a dynamic range of about 100…”

First at all, the expression “dynamic range” has to be replace by “signal-to-noise
ratio” (SNR) – sorry for my fault. It makes only sense for a detector not for an
image itself.

Secondly, the value 100 was estimated by 10,000 incident electrons per pixel –
resulting in a noise 100 = sqrt(10,000). Therefore the SNR of a captured image is
about 100 (=10,000/100) – assuming a “normal” specimen (e.g. weak-phase object).
But please don’t pin me down at numbers, basically I wanted to express that for
normal (bright-field) applications an image detector with a dynamic range of a few
hundred is sufficient. Actually a photo plate has “only” such a dynamic range.
But to make it clear: A SNR of 100 is an excellent value – a rough rule of thumb
says the SNR should be at least higher than 5 (so-called “Rose criteria”). I don’t
dare to give a reference of this old and famous work from 1946 :-)

I feel that this question may arise: “But why are you guys selling CCD cameras
which such a high dynamic range if normal users don’t need it?”
A: Normally, CCD cameras are designed to detect single electrons – especially for
low-dose/cryo applications. If you want to capture a “high-dose” image with 10,000
electrons/pixel, then the huge dynamic range is needed. This means CCD cameras are
able to handle both applications.

One remark regarding Bill (and Mike): I agree with Mike, the non-linearity of CCD
cameras is normally less than 1%.

Best wishes,
Ingo

+++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Dr. Ingo Daberkow
Tietz Video and Image Processing Systems GmbH
Eremitenweg 1
D-82131 Gauting, Germany
Tel: +49-89-8506567
FAX: +49-89-8509488
Internet: http://www.tvips.com/
Email: ingo.daberkow-at-tvips.com
+++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++




Sergey Ryazantsev wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Ingo
} Thanks for comments. I absolutely agree with folks who mentioned extremely
} wide dynamic range for image plates. I think there is some misunderstood
} happening: In order to get data from image plate, you have to scan it and
} record it as a digital image. The quality of scanners is different, some
} of them could not provide 20 bit gray scale. When I pointed, that dynamic
} range is the same for the images from CCD and IP I mean that the images
} were recorded in 12-16 bit format, which is very usual for nova days. 20
} bit format is non-readable by most software and therefore useless for such
} simple person like me, who is still using Photoshop for most jobs. So,
} from this point of view, we have practically the same dynamic range of the
} final digital image. I am sorry if my posting confuse somebody. Image
} plate technology widely used in X-ray crystallography and as a substitution
} for X-ray film in molecular biology applications for decades. No question
} about that: this is a great technology. BUT, my point, is that current
} realization of this technology in EM does not have serious advantages over
} normal film/digital camera combination. AND: keep in mind, how long you
} will scan to get those 20 bits dynamic range... Thanks again for the very
} good point. Sergey
}
} At 01:13 PM 6/5/02, you wrote:
} } Dear Sergey,
} }
} } Just another remark regarding the dynamic range. The imaging plates have a
} } much
} } higher dynamic range than other electron detectors (up to 20 bit after several
} } scans). But please notice the dynamic range is defined as: (maximum signal/RMS
} } noise) ­ this is not the digitization level of the analog-to-digital
} } converter.
} } This extreme high dynamic range is only necessary for the acquisition of
} } diffraction patterns. A “normal” bright-field image has typically a dynamic
} } range of about 100, which is mainly limited by the noise of the electrons
} } (shot-noise). This means, for such applications CCD cameras have the advantage
} } of a real-time acquisition (WYSIWYG) and furthermore the camera can be
} } used for
} } checking or controlling of the TEM parameters (defocus, astigmatism, etc.)
} }
} } Personally I have found the literature list of Wharton very useful ­ which
} } enables the interesting reader to collect more specific information.
} }
} } Best wishes,
} } Ingo
} }
} } +++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
} } Dr. Ingo Daberkow
} } Tietz Video and Image Processing Systems GmbH
} } Eremitenweg 1
} } D-82131 Gauting, Germany
} } Tel: +49-89-8506567
} } FAX: +49-89-8509488
} } Internet: http://www.tvips.com/
} } Email: ingo.daberkow-at-tvips.com
} } +++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
} }
} } ,
} }
} }
} }
} } "Sinkler, Wharton" wrote:
} }
} } } ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------.
} } }
} } } OK, Truce Sergey!
} } }
} } } Your point is well taken, and yes, design considerations and
} } convenience may
} } } overwhelm a good MTF or linearity.
} } }
} } } However, I think you are mistaken about dynamic range being similar for
} } } Imaging Plates and CCD's. That being said, I have gone back to look at the
} } } papers I mentioned and I'm not able to find a clear statement of what I
} } } mean.
} } }
} } } I believe (maybe someone will set us straight on this) that the saturation
} } } of the IP is considerably higher than the CCD. By 'saturation' I mean the
} } } dose (electrons/unit area) at which there is significant deviation from
} } } linearity.
} } }
} } } The reason saturation level is important (and why the IP is great for
} } } diffraction) is that signal/noise at low electron doses is dominated by
} } shot
} } } noise. So if you are trying to record a diffraction pattern with a CCD
} } with
} } } a single exposure without saturating the CCD at the strong spots, the weak
} } } reflections will be noisy simply because relatively few electrons are
} } } arriving there (statistical or 'shot' noise). With a higher saturation (I
} } } believe the case for IP), you can expose longer and get better statistics
} } } overall. (Of course you can also record multiple exposures with the CCD at
} } } different times and merge the data).
} } }
} } } Regards,
} } } Wharton
} } }
} } } } -----Original Message-----
} } } } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } } } Sent: Tuesday, June 04, 2002 7:42 PM
} } } } To: Microscopy-at-sparc5.microscopy.com
} } } } Subject: RE: TEM imaging plate technology
} } } }
} } } } ------------------------------------------------------------------------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } -----------------------------------------------------------------------.
} } } }
} } } }
} } } } Dear Wharton
} } } }
} } } } Are you seriously think, people will read that very specific papers? Do
} } } } you think, so many of us is ready to go through all this math etc? I
} } } } agree,
} } } } it's very important to understand the scientific foundation for
} } device you
} } } }
} } } } are going to use. But, in practice you have deal not with such nice
} } } } things
} } } } as PSF (point spread function) or MTF. The camera with very nice PSF may
} } } } have terrible design and not suitable for use. Company, who manufactured
} } } } that baby may not respond on your telephone calls and finally you may
} } } } figure out that it does not exists anymore. I do read scientific
} } } } journals
} } } } and it's extremely easy to find informations on the Internet nova days,
} } } } ListServer to me is a place where we have chance to share our personal
} } } } experience (bad or good), which mostly is not published and not
} } indexed by
} } } }
} } } } Medline/CC whatever. Anyway, thank you for the reference, I enjoyed
} } } } reading it. Sergey
} } } }
} } } } At 01:41 PM 6/4/02, you wrote:
} } } }
} } } } } Sergey et al,
} } } } }
} } } } } I'd just like to point out that instead of relying on uncertain
} } } } } recollections, there is some good work in the literature which discusses
} } } } } both slow-scan CCD's and imaging plates. Here's a start:
} } } } }
} } } } } J. M. Zuo "Electron detection characteristics of slow-scan CCD camera"
} } } } } Ultramicroscopy 66 (1996) 21-33.
} } } } }
} } } } } J. M. Zuo, M. R. McCartney and J. C. H. Spence "Performance of imaging
} } } } } plates for electron recording" Ultramicroscopy 66 (1996) 35-47
} } } } }
} } } } } G. Y. Fan and M. H. Ellisman "Digital imaging in transmission electron
} } } } } microscopy" Journal of Microscopy 200 (2000) 1-13.
} } } } }
} } } } } Although the technology is continuously evolving, the basic points are
} } } } still
} } } } } valid.
} } } } }
} } } } } Regards,
} } } } } Wharton
} } } } }
} } } } } *************************************************
} } } } } Wharton Sinkler, Ph.D.
} } } } } UOP LLC
} } } } } 25 E. Algonquin Rd.
} } } } } Des Plaines, IL 60017-5017
} } } } } 847-391-3878
} } } } }
} } } } }
} } } } } } -----Original Message-----
} } } } } } From: Sergey Ryazantsev [SMTP:sryazant-at-ucla.edu]
} } } } } } Sent: Tuesday, June 04, 2002 3:36 PM
} } } } } } To: microscopy-at-sparc5.microscopy.com
} } } } } } Subject: Re: TEM imaging plate technology
} } } } } }
} } } } } }
} } } } ------------------------------------------------------------------------
} } } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } } } America
} } } } } } To Subscribe/Unsubscribe -- Send Email to
} } } } ListServer-at-MSA.Microscopy.Com
} } } } } } On-Line Help
} } } } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } } }
} } } } -----------------------------------------------------------------------.
} } } } } }
} } } } } }
} } } } } } Dear Donald
} } } } } }
} } } } } } Before I bought the digital camera, I was looking on that image plate
} } } } } } technology. It seems to me, it does not have serious advantage for
} } } } most
} } } } } } of
} } } } } } us. I would like to concentrate my findings in a few sentences.
} } } } } }
} } } } } } Image plate PRO:
} } } } } } - has more pixels that most digital cameras (some top-end cameras
} } } } } } HAS similar amount).
} } } } } } - has dynamic range same as digital cameras 12-16 bit.
} } } } } } - it's more sensitive than film.
} } } } } } - very good for electron diffraction experiments!!! (only one real
} } } } plus to
} } } } } } me).
} } } } } } - manufacturers claimed that pixel resolution on the plate is
} } } } comparable
} } } } } } with film (it's difficult to proof because you have to scan film to
} } } } } } compare, so it'll depend from the film scanner).
} } } } } }
} } } } } } Image plate CONTRA:
} } } } } } Imitate the film procedure - you have to perform all procedures
} } as for
} } } } } } film, load/upload plates into cassettes (in the dark), change the
} } } } magazine
} } } } } }
} } } } } } (wait for vacuum), load plates into the scanner (in the dark I
} } } } believe),
} } } } } } wait for scanning - 2 (or more, don't remember, up to 5 at full
} } } } resolution
} } } } } }
} } } } } } I believe) min etc. So, it does not eliminate the dark-room,
} } scanning
} } } } is
} } } } } } slow and then you have to process/save/print data. Time consuming.
} } } } You
} } } } } } have all disadvantages the classical film use: plate may be scratched
} } } } } } during loading/uploading, deformed, lost, dropped on floor with
} } } } valuable
} } } } } } image etc. One plate is $100 I believe. No practical use in most
} } } } } } biological applications I believe. I am not warranty that all my
} } } } } } information is current, I was shopping for image plate system a few
} } } } years
} } } } } } ago and my comments represent the situation of that time. Sorry,
} } } } } } manufacturing gays. Sergey
} } } } } }
} } } } } } At 08:07 AM 6/4/02 -0500, you wrote:
} } } } } }
} } } } } ------------------------------------------------------------------------
} } } } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } } } America
} } } } } } } To Subscribe/Unsubscribe -- Send Email to
} } } } ListServer-at-MSA.Microscopy.Com
} } } } } } } On-Line Help
} } } } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } } }
} } } } } -----------------------------------------------------------------------.
} } } } } } }
} } } } } } }
} } } } } } }
} } } } } } }
} } } } } } } Dear EM Netters,
} } } } } } }
} } } } } } } Does anyoun out there have experience with imaging plate technology
} } } } (i.e.
} } } } } } } Ditabis)? I
} } } } } } } would like to hear pros and cons concidering this technology. How
} } } } does
} } } } } } it
} } } } } } } compare to
} } } } } } } digital CCD camera systems (cost, time, supplies, reliability. We
} } } } are
} } } } } } } looking into possibly
} } } } } } } retrofitting our JEOL 100CX with an off-line system.
} } } } } } }
} } } } } } }
} } } } } } } Thanks in advance
} } } } } } }
} } } } } } }
} } } } } } } Donald G. Awbrey, HT (ASCP), QIHC
} } } } } } } Electron Microscopy / Image Analysis
} } } } } } } 817-878-5647
} } } } } } } donaldawbrey-at-texashealth.org
} } } } } } }
} } } } } }
} } } } } } ------------------------------------------------------
} } } } } }
} } } } } } Sergey Ryazantsev, Ph.D.
} } } } } } Electron Microscopy
} } } } } } Department of Biological Chemistry, School of Medicine
} } } } } } University of California, Los Angeles
} } } } } } Box 951737
} } } } } } Los Angeles, CA 90095-1737
} } } } } }
} } } } } } (310) 825-1144 (office)
} } } } } } Pager: (310) 845-0248
} } } } } } FAX: (310) 206-5272 (departmental)
} } } } } } mailto:sryazant-at-ucla.edu
} } } } } }
} } } }
} } } } _____________________________________
} } } }
} } } } Sergey Ryazantsev Ph. D.
} } } } Electron Microscopy
} } } } UCLA School of Medicine
} } } } Department of Biological Chemistry
} } } } Box 951737
} } } } Los Angeles, CA 90095-1737
} } } }
} } } } Phone: (310) 825-1144
} } } } FAX (departmental): (310) 206-5272
} } } } mailto:sryazant-at-ucla.edu
} } } }
} } } }
} } } }
}
} _____________________________________
}
} Sergey Ryazantsev Ph. D.
} Electron Microscopy
} UCLA School of Medicine
} Department of Biological Chemistry
} Box 951737
} Los Angeles, CA 90095-1737
}
} Phone: (310) 825-1144
} FAX (departmental): (310) 206-5272
} mailto:sryazant-at-ucla.edu



From daemon Thu Jun 6 10:23:02 2002



From: Battjes, Kevin :      battjes-at-impactanalytical.com
Date: Thu, 6 Jun 2002 11:16:14 -0400
Subject: Chillers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lou,
We have had Haskris chillers, both water cooled and air cooled for many
years. They run 24/7 very faithfully. I would buy another one.

Disclaimer: Haskris has never bought me lunch or coffee or even a candy
bar! I have no commercial interest in the sales of their equipment.


Kevin Battjes
Impact Analytical Voice 989-832-5555, ext 556
Michigan Molecular Institute Fax 989-832-5560
1910 W. St Andrews Road e-mail: battjes-at-mmi.org
Midland MI 48640 battjes-at-impactanalytical.com




-----Original Message-----
} From: Lou bustillos [mailto:lbustillos-at-amalab.com]
Sent: Thursday, June 06, 2002 9:53 AM
To: Microscopy-at-sparc5.microscopy.com


Hello,

First I would like to say thank you to everyone that helped me with my last
email about purchasing a carbon coater. Hopefully installation will be in
the next two weeks.

My new question is about chillers for a JEOL 100CX II. The state that we
are located in is beginning to have a drought and we want to do our part by
changing our water chillers to air cooled chillers. I would like to know
your experience with switching to an air cooled system. Also, what
manufacture would you recommend? I have two 100CX II and they will have
there own separate units. Thank you for your help.



From daemon Thu Jun 6 12:56:36 2002



From: Anthony J. Garratt-Reed :      tonygr-at-mit.edu
Date: Thu, 06 Jun 2002 13:41:08 -0400
Subject: Re: Chillers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lou-

Air cooled chillers work just fine. They produce lots of heat, so they
will need good ventilation in the summer. However, they cannot be in a
position where they would freeze in winter if switched off for any reason
(e.g. power failure). They also tend to be quite noisy.

We have systems from several manufacturers. They all have been good. I
would shop around for the usual best combination of performance, features
and price appropriate for your application.

It doesn't apply to you, but a user switching from using the public water
supply for direct cooling will have been using a system with extremely good
short-term temperature stability. A simple chiller does not have that -
typically it will vary by about 2şC over its cycle time - i.e. a few
minutes. This will usually be very clear in the image drift! There are a
number of methods of improving the short-term temperature stability which
can be discussed with the chiller manufacturers.

We have always used recirculating chillers, using water-cooled models. We
switched several years ago from the city water as the coolant for the
condenser to the chilled water provided by the facilities people for the
air conditioning. After a few teething troubles (which I would be glad to
share with anyone interested) our system now works reliably, doesn't waste
water, doesn't dump heat into the rooms, isn't too noisy, and didn't
require us to buy 6 new air-cooled chillers! Of course not everyone will
have that option.

Tony.



} Hello,
}
} First I would like to say thank you to everyone that helped me with my last
} email about purchasing a carbon coater. Hopefully installation will be in
} the next two weeks.
}
} My new question is about chillers for a JEOL 100CX II. The state that we
} are located in is beginning to have a drought and we want to do our part by
} changing our water chillers to air cooled chillers. I would like to know
} your experience with switching to an air cooled system. Also, what
} manufacture would you recommend? I have two 100CX II and they will have
} there own separate units. Thank you for your help.


** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6478
**




From daemon Thu Jun 6 13:53:13 2002



From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Thu, 6 Jun 2002 14:49:07 -0400
Subject: Thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nestor:

I really enjou the listserver, and think I get a great deal of good
stuff from it. I think you do a great job managing it, and thank you
immensely for all the effort you put into doing so.

Best regards,
Wil Bigelow
--
Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
3062 Dow Bldg.; 2300 Hayward St.
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:734-763-4788; Ph:734-662-5237


From daemon Thu Jun 6 13:53:15 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Thu, 06 Jun 2002 11:59:14 -0700
Subject: RE: virus alert

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fred!

Thanks for your message. I really appreciate your respond. It's so nice
of you. Of coarse, I never ever intend to think that somebody from
ListServer sent to me virus. It was sort of joke (my 'humor' was a little
bit heavy, I guess). As a matter of fact, I survived a few complete (!)
computer crushes with following HD reformat. In all that cases I got
viruses from friends (lost attention). So, applying this idea to our
situation: everyone on ListServer is my friends! And it's very
true. ListServer become to me a place where friends are situated. I am
really sorry if my last posting somehow hurt our ListServerers. I really
enjoy being a part of this society and thank you so much for attention.
Sergey


} } ----------
} } From: Sergey Ryazantsev
} } Sent: Wednesday, June 5, 2002 3:12 PM
} } To: microscopy-at-sparc5.microscopy.com
} } Subject: virus alert
} } Importance: High
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } I just got message with subject "Subject: To call in the service
} } representative." from microscopy-at-sparc5.microscopy.com with attachment
} } aa_160x100[1].exe
} }
} } This message contains some virus. Unfortunately my Norton Antivirus
} } eliminate it so quickly and I have no chance to find what virus it was.
} } Thanks for such nice gift. May be it's sort of sign, I am talking too
} } much
} } on ListServer? Sergey
} } ------------------------------------------------------
} }
} } Sergey Ryazantsev, Ph.D.
} } Electron Microscopy
} } Department of Biological Chemistry, School of Medicine
} } University of California, Los Angeles
} } Box 951737
} } Los Angeles, CA 90095-1737
} }
} } (310) 825-1144 (office)
} } Pager: (310) 845-0248
} } FAX: (310) 206-5272 (departmental)
} } mailto:sryazant-at-ucla.edu
} }
} }
} }

------------------------------------------------------

Sergey Ryazantsev, Ph.D.
Electron Microscopy
Department of Biological Chemistry, School of Medicine
University of California, Los Angeles
Box 951737
Los Angeles, CA 90095-1737

(310) 825-1144 (office)
Pager: (310) 845-0248
FAX: (310) 206-5272 (departmental)
mailto:sryazant-at-ucla.edu



From daemon Thu Jun 6 15:40:53 2002



From: Allan Mitchell :      allan.mitchell-at-stonebow.otago.ac.nz
Date: Fri, 7 Jun 2002 08:31:30 +1200
Subject: re-embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Hong

Dug out and dusted off these references for re-embedding last night.


1. Method of re-embedding tissue for electron microscopy. Stain
Technology Vol49, page 118 to 119. (1974) Bauman and Mendall

Done with Spurrs.

- Trimmed away excess plastic under a stereo microscope with razor blade
- tissue put in vial of Propylene Oxide and sonicated for 1 hour
- excess plastic (now soft and pliable) again trimmed away
- tissue infiltrated with 1:1 resin/ propylene oxide with gental
agitation for 18 hours at RT
- 48 hours infiltration in full strength Spurrs, again gentle agitation at RT
- re-embed and polymerise

We have tried this with tissue blocks and it worked for us. Don't
know how vibratome slices would stand up to it.



2. Rescuing poorly embeded tissue for electron microscopy: a new and
simple technique for re-embedding. Stain Technology Vol50, page 209
to 211. (1975) McNelly and Hinds

Done with Araldite 502.

- place 10µm (or thin as can get) slices placed in disposable
aluminum weighing dishes
- cover sections with distilled water and heat on hot plate until
water evaporated, leaving sections stuck on the bottom.
- fill dishes with resin mixture and place in a dessicator over night
(doesn't mention vaccum level or use of a desicant)
- then transfer to an oven at 60oC for 48 hours.
- remove polymerised block cut out area of interest, remount and away you go

I haven't tried this one so no comment on its success.


3. I don't have the reference for this one but it is mentioned in
Principles and Techniques of Electron Microscopy, MA Hayat, Third
Edition, page 133

Method of Ogura and Oda, 1973.

-trimed excess plastic away, soaked in 100% ethanol saturated with
KOH for several hours. The ethanol/KOH must be matured overnight
before use (becomes dark brown).

- after solubilisation of the resin soak in 100% ethanol, followed by
propylene oxide, then re-embedded in resin.

Interestingly McNelly and Hinds and say in their paper they tried
this and it didn't work for them. I haven't tried it either so can't
comment.

As I mentioned above, the only one I have tried is 1. above and it
worked for us.

Good luck,

Allan

PS. have you spotted any legs under any bar tables lately that match
or better Doug's yet ?
--
-------------------------------------------------
Allan Mitchell
Technical Manager
Otago Centre for Electron Microscopy
C/-Department of Anatomy and Structural Biology
School of Medical Sciences
P.O. Box 913
Dunedin
New Zealand

Phone (03) 479 5642 or 479 7301
Fax (03) 479 7254

Unit: http://www.otago.ac.nz/anatomy/emunit/
Department: http://anatomy.otago.ac.nz/



"Life is a gift, don't waste it"


From daemon Thu Jun 6 17:44:52 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Thu, 06 Jun 2002 15:36:13 -0700
Subject: RE: TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Gene

I am using VNC software, which is stands for Virtual Network
Computing. It's freeware, look at http://www.uk.research.att.com/vnc/

VNC is developed to transmit ANY information via network, not necessarily
images from digital camera. Actually, it transmits the whole screen in the
real time (depends from your network speed). So, If you are working with
digital camera, it'll transmit everything you doing/seeing. It has a
setting, when remote mouse/keyboard is activated. In this case remote
operator may manipulate your computer/camera/software. If microscope
controlled by computer, you may control the whole things. The reasons, I
am using this software rather than NetMeeting or something like that are:
VNC is very compact and should be installed on the 'transmitter computer'
only (VNC-server, transmitter). It may works as a 'service' or as a
standard application under WinNT/2000 (it does not work under Win95/8). All
other computers - recipients, should run "VNC viewer" - 34 kB .exe file
only (there is no installation, just run). You could run it even from
diskette on any computer (OS limitation). In order to use VNC, you don't
have to use the 'outside server' as it happening with NetMeeting, ICQ etc:
You connected to MS server and it transmit your session to another
'user'. VNC transmits to 'everyone' inside the network, and you may get
this transmission if you run VNC-viewer and provide correct password
(password is not transmitted via network). If you permit transmission to
Internet - everyone on the Internet will have chance to catch this
transmission, this is a disadvantage. For 'Internet' session, I've sent
by E.mail VNC-viewer file, preconfigured for listening my transmission and
password by phone. Remote user should double-click on the VNC-viewer.exe
file and enter password. That's it. Only one things here: remote user
should not be such picky/panic as me about viruses in .exe files.

The disadvantage of VNC is: It designed for local network and therefore has
very limited security (I belive, it uses standard TCP/IP protocol), so the
computer running 'VNC server, read - transmitter' is vulnerable if
available from the Internet. The best way to use VNC is to use it inside
the local network, protected from outside by firewall and never run it on
your 'departmental server'. I am not running VNC permanently, instead, I
set 'session' when my computer is running this software. For each session
I provide a new password. If you have good network administrator, s/he
could probably advise you, how to use this software safely in your
particular case.

I hope it could help. Good luck. Sergey

At 06:38 AM 6/6/02, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Thu Jun 6 23:38:30 2002



From: John Foust :      jfoust-at-threedee.com
Date: Thu, 06 Jun 2002 23:24:07 -0500
Subject: RE: TEM-Digital Camera Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 03:36 PM 6/6/2002 -0700, Sergey Ryazantsev wrote:
} I am using VNC software, which is stands for Virtual Network Computing. It's freeware, look at http://www.uk.research.att.com/vnc/
} It may works as a 'service' or as a standard application under WinNT/2000 (it does not work under Win95/8).

VNC server and viewer works fine under Windows 98.
Also see http://www.tightvnc.com/ . It's also multi-platform,
so someone on a Mac can watch a PC. I've used the VNC viewer
on palmtops and even a Palm, I think.

- John



From daemon Fri Jun 7 02:43:18 2002



From: Chris Jeffree :      c.jeffree-at-ed.ac.uk
Date: Fri, 7 Jun 2002 08:38:17 +0100
Subject: Hitachi / EDAX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Is anyone out there using EDAX Phoenix on a Hitachi 4700?
If so, I would very much like to make contact, offline

Dr. Chris Jeffree
University of Edinburgh
Biological Sciences EM Facility
EDINBURGH EH9 3JN UK

Inveresk Cottage
26, Carberry Road
Inveresk
Musselburgh
Midlothian
EH21 8PR
Tel: +44 131 665 6062
FAX +44 131 653 6248
Mobile 07710 585 401



From daemon Fri Jun 7 04:35:02 2002



From: PHYSIOL4-at-AKAD.SUN.AC.ZA
Date: Fri, 7 Jun 2002 11:26:26 +0200
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe


"When life hands you a lemon ... bring out the tequila and salt."


From daemon Fri Jun 7 09:53:39 2002



From: Jeannette Taylor :      jvtaylo-at-emory.edu
Date: Fri, 07 Jun 2002 10:50:32 -0400
Subject: negative stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have just had an interesting problem show up, or not show up as it
were. One of the students here put her negatively stained grids into
the TEM to re-examin after a couple of months and there appeared to be
no stain. She used K-PTA at pH 2 for 4 minutes and got an a excess of
stain initially. I have not re-viewed negatively stained grids once I
have photographed them so I don't know what might have happened.
Neither have I heard of this before.
Does the List have experience and comments on this? Please do.

Thanks.
Jeannette Taylor
IM&MF/ Emory University
jvtaylo-at-emory.edu



From daemon Fri Jun 7 10:24:12 2002



From: David Hall :      hall-at-aecom.yu.edu
Date: Fri, 7 Jun 2002 11:17:09 -0400
Subject: LKB microtome manual

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for the operations manual for an LKB Ultratome III.
Could someone contact me offline and FAX or mail us a copy? Thanks
so much.
--
David H. Hall, Ph.D.
Center for C. elegans Anatomy
Department of Neuroscience
Albert Einstein College of Medicine
1410 Pelham Parkway
Bronx, NY 10461

www.wormatlas.org
www.aecom.yu.edu/wormem

phone 718 430-2195
fax 718 430-8821


From daemon Fri Jun 7 12:58:56 2002



From: Michael L. Boucher :      mboucher-at-tc.umn.edu
Date: Fri, 07 Jun 2002 12:52:19 -0500
Subject: Supplier of Robinson Chamber View?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a Robinson Chamber View camera that we need to modify to fit a new
SEM. Does anyone know where I can contact the company or U.S. distributor?
Thanks for any replies.
Regards,
Mike
********************************************************************
Michael L. Boucher Sr. mboucher-at-tc.umn.edu
Lab Manager Rm 55 Office Ph 612-624-6590
I.T. Characterization Facility Rm 12 Main Off Ph 612-626-7594
University of MN Fax 612-625-5368
12 Shepherd Labs
100 Union Street S.E.
Minneapolis, MN 55455 http://resolution.umn.edu
********************************************************************








From daemon Fri Jun 7 13:26:58 2002



From: Bill & Sue Tivol :      wtivol-at-earthlink.net
Date: Fri, 07 Jun 2002 11:18:33 -0400
Subject: Re: Chillers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


on 6/6/02 1:41 PM, Anthony J. Garratt-Reed at tonygr-at-mit.edu wrote:
}
} It doesn't apply to you, but a user switching from using the public water
} supply for direct cooling will have been using a system with extremely good
} short-term temperature stability. A simple chiller does not have that -
} typically it will vary by about 2şC over its cycle time - i.e. a few
} minutes. This will usually be very clear in the image drift! There are a
} number of methods of improving the short-term temperature stability which
} can be discussed with the chiller manufacturers.
}
Dear Tony & Lou,
Haskris, and, I'm sure, other manufacturers, makes a hot-gas bypass
attachment which provides an order of magnitude better temperature
stability. We tracked the cyclic drift from the temp changes before
ordering the bypass unit, and, when we tested for drift after the unit was
installed, the drift was negligable. BTW, the drift test suggested in the
manual--take pictures every half hour--did not catch the cyclic drift; we
had to watch for several minutes to see the problem.
Yours,
Bill Tivol



From daemon Fri Jun 7 13:53:18 2002



From: Russell Spear :      rzs-at-plantpath.wisc.edu
Date: Fri, 7 Jun 2002 13:43:28 CST
Subject: vernier scales

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a source of adhesive vernier scales that could be
attached to a microscope stage.

Thanks

Russ Spear
Russell N. Spear
Sr. Research Specialist
Dept. of Plant Pathology
Univ. of Wisconsin-Madison

RZS-at-plantpath.wisc.edu
Phone 608 263-2093
Fax 608 263-2626


From daemon Fri Jun 7 14:25:31 2002



From: Xianglin Li :      Xianglin_Li-at-student.uml.edu
Date: Fri, 07 Jun 2002 15:16:16 -0400
Subject: Asking for SEM machine

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friends,

Our center is looking for a "new" SEM machine (with good condition and
resolution, within last 10 years). We want this machine could do EDX, X-
ray mapping, and with LaB6 as the filament. Of course, the cheaper, the
better.

Is anybody have a good deal for that? We will appreciate that!

Regards,

Xianglin Li

Center for Advanced Material
Department of Chemical Engineering
University of Massachusetts, Lowell
Xianglin_Li-at-student.uml.edu
Tel: 978-934-3411




From daemon Fri Jun 7 15:20:15 2002



From: Johnson, Bradley R :      Bradley.Johnson-at-pnl.gov
Date: Fri, 07 Jun 2002 13:12:49 -0700
Subject: RE: Supplier of Robinson Chamber View?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Mike,
Try contacting Bob Rusica:


Robert Ruscica
ETP-USA
Tel:(916)797-6199
Fax:(916)797-6304
email:ruscica-at-etp-usa.com
www.etp-usa.com


-Brad


----------
From: Michael L. Boucher
Sent: Friday, June 7, 2002 10:52 AM
To: Microscopy-at-sparc5.microscopy.com
Subject: Supplier of Robinson Chamber View?


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

-----------------------------------------------------------------------.


We have a Robinson Chamber View camera that we need to modify to fit
a new
SEM. Does anyone know where I can contact the company or U.S.
distributor?
Thanks for any replies.
Regards,
Mike
********************************************************************
Michael L. Boucher Sr. mboucher-at-tc.umn.edu
Lab Manager Rm 55 Office Ph 612-624-6590
I.T. Characterization Facility Rm 12 Main Off Ph 612-626-7594
University of MN Fax 612-625-5368
12 Shepherd Labs
100 Union Street S.E.
Minneapolis, MN 55455 http://resolution.umn.edu
********************************************************************










From daemon Fri Jun 7 22:50:17 2002



From: Gary Gaugler :      gary-at-gaugler.com
Date: Fri, 07 Jun 2002 20:42:04 -0700
Subject: Glow Plug SEM analysis inquiry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am working on a rather fun subject of a model
engine glow plug. I have put some SEM pix at

http://www.microtechnics.com/glow.htm

which will pull up a PDF catalog of SEM images.
I think that the key images are the BSE ones,
including 10, 2, and 3. What I do not know is
what is going on in these pix. I tend to see a
phase change, but the metal coil wire is at room
temperature.

If anyone has any ideas about what is seen in
these pix, please reply off-line with your comments
and attribution data. As I am planning on writing
an article about this topic, I will include relevant
feedback and will attribution specifics.

One item of interest is that this particular plug was
relatively new (about 45 minutes of running from
totally new). It faltered and was replaced by a new
plug which solved the problem. so, why does this
happen and what exactly is happening to the plugs
to degrade them? Beats me.

gary g.



From daemon Sat Jun 8 00:43:33 2002



From: Allen Sampson :      ars-at-sem.com
Date: Sat, 8 Jun 2002 00:37:24 -0700
Subject: RE: Glow Plug SEM analysis inquiry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Any idea what the wire composition is? How about the fuel composition and
fuel/air mixture being used? What imaging was used for 8 & 9 (is 9 BSE)?
Were 8 & 9 taken at the upper coil connection point? If so, that bond
looks suspicious, unless the rest that's not shown shows a better bond.
Anything more specific on failure mode than 'faltered'?

Allen R. Sampson
Advanced Research Systems
317 North 4th. Street
St. Charles, Illinois 60174

phone (630) 513-7093 fax (630) 513-7092 http://www.sem.com


On Friday, June 07, 2002 8:42 PM, Gary Gaugler [SMTP:gary-at-gaugler.com]
wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I am working on a rather fun subject of a model
} engine glow plug. I have put some SEM pix at
}
} http://www.microtechnics.com/glow.htm
}
} which will pull up a PDF catalog of SEM images.
} I think that the key images are the BSE ones,
} including 10, 2, and 3. What I do not know is
} what is going on in these pix. I tend to see a
} phase change, but the metal coil wire is at room
} temperature.
}
} If anyone has any ideas about what is seen in
} these pix, please reply off-line with your comments
} and attribution data. As I am planning on writing
} an article about this topic, I will include relevant
} feedback and will attribution specifics.
}
} One item of interest is that this particular plug was
} relatively new (about 45 minutes of running from
} totally new). It faltered and was replaced by a new
} plug which solved the problem. so, why does this
} happen and what exactly is happening to the plugs
} to degrade them? Beats me.
}
} gary g.
}
}
}
}



From daemon Sat Jun 8 01:04:43 2002



From: paul r hazelton, PhD :      Paul_Hazelton-at-umanitoba.ca
Date: Sat, 08 Jun 2002 01:00:49 -0500
Subject: Re: negative stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


jeannette

well, it certainly is an interesting protocol. but then, i know one
person who insists that staining for more than 10 seconds gives positive
staining, not negative staining, while another collaborator in a
recently submitted review suggested that he uses 5 minutes staining
time. in my experience, most staining effect is seen in the first 30-60
seconds.

what is intersting is the pH of the stain. i have seen Uranyl stains
used at low pH, but never PTA. i know that use of PTA has been reported
as low as pH 4.0, but i have only seen uncited references to this, so i
do not know what the actual studies showed. my memory is also that the
natural pH of PTA is already about 2.0, so you must not have done a lot
of adjusting. i have personally done some work at pH 6.0 and pH 8.0,
but most is at 7.0.

i know KOH can be used for pH adjustment, but it is also associated with
the degradation of virions. look at old work by Hoyle, early '60's,
where they eventually twigged that when they used KOH to neutralize PTA
the viral envelopes were degraded, but when they used NaOH the particles
did not degrade unless other treatment, which was intended to degrade
the particle, was used. the use of KOH is probably the source of the
arguement that PTA degrades the particles.

when we use NaOH we do not see particle degradation or loss of
significant loss in stain density over short time the particles are
clearly readable, with no apparant degradation over several weeks. the
biggest degradation seems to occur after being put in the scope and
exposed to vacuum and beam, and that is not major in our experience.

as far as stability of stain on the grid during prolonged storage - i
have set aside grids prepared from clinical samples as training grids
and looked at them 2-3 years later and seen clearly defined virus. the
only problem seems to occur when the grid is subjected to prolonged
exposure to beam and vacuum, which sometimes occurs because frequent
reference to the these grids by students or technicians who are in
training.

to the point, wednesday i had reason to look at a grid with subviral
particles that was prepared by a student in november. the staining was
still good, and i was able to particle count the sample. the observed
distribution of particle types, and subviral components, was consistent
with previous experiments, as was the total number of particles
observed. while several micrographs were taken for archival purpose,
they will probably never be printed. however, those negatives were
scanned through the enlarger and do look good.

in short, the problem should not be time of storage. but i would look
seriously at the use of NaOH instead of KOH for neutralization. i would
also look at higher pH, unless there is an isoelectric point issue that
you must address.

now, i must admit that i'm interested in what sara miller has to say on
the issue.

paul hazelton





From daemon Sat Jun 8 17:02:18 2002



From: Rooting :      rooting-at-hortus.com
Date: Sat, 8 Jun 2002 16:42:44 -0500
Subject: vernier scales

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a source of adhesive vernier scales that could be
attached to a microscope stage.

Thanks

Russ Spear
Russell N. Spear
Sr. Research Specialist
Dept. of Plant Pathology
Univ. of Wisconsin-Madison




McmasterCarr has some crude ones in their "adhesive backed rules" section

http://www.mcmaster.com

They are pretty inexpensive and Mcmaster does not have minimum orders and
low actual shipping chg and takes credit cards. They usually ship same day.

regards
MRK


From daemon Sat Jun 8 18:38:29 2002



From: Qian-Chun Yu, MB, Ph.D. :      qcyu-at-mail.med.upenn.edu
Date: Sun, 09 Jun 2002 09:12:17 -0400
Subject: just testing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Here is a source of small amounts of Canadian Balsam at reasonable prices.
http://studioproducts.com/catalog/balsams/canad1.html
It is from an artist supply house but it is reported to be very good quality
and suitable for slide making.

I am collecting sources of difficult to find items useful information and
posting it on a page at:
http://www.couger.com/microscope/links/gclinks.html

Any sources that any of you have for supplies would be appreciated so I can
add them to this list.

Gordon
----- Original Message -----
} From: "Rooting" {rooting-at-hortus.com}
To: "Microscope" {Microscopy-at-sparc5.microscopy.com} ; "Microscope list,"
{microscopes-at-yahoogroups.com} ; "Microscope Confocal,"
{confocal-at-listserv.acsu.buffalo.edu}
Sent: Saturday, June 08, 2002 5:40 PM


this is just to test that my message will be able to reach the community.
please disregard any information that I may put in here. thanks!

QCY



From daemon Mon Jun 10 00:16:52 2002



From: Gary Gaugler :      gary-at-gaugler.com
Date: Sun, 09 Jun 2002 22:02:37 -0700
Subject: Looking for EDX detector

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


It's becoming increasingly clear to me that I could
greatly benefit from having an EDX system on my
FESEM. The most cost advantageous, and feature-based
solution seems to be to find a surplus detector and
mate it with electronics and software from IXRF.

If you have a homeless detector (Si(Li) or other type)
that uses LN2 or cryo cooling, please contact me off-line
and let me know the details of what you have....and
cost of transfer.

I'm hoping to mate the detector to an Amray 1910FESEM.
It would use the left rear port (looking at the chamber
from the front).

gary g.



From daemon Mon Jun 10 06:16:43 2002



From: Oparowski, Joseph :      Joseph_Oparowski-at-bose.com
Date: Mon, 10 Jun 2002 07:09:00 -0400
Subject: Supplier of Robinson Chamber View?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mike,

You can contact the US rep at;

ETP-USA
Electron Detectors Inc.
4734 Tenbury Lane
Rocklin, CA 95677
(916) 797-6199
Bob Ruscica

Good luck.


Joseph M. Oparowski
Materials Science Research - Consulting and Failure Analysis
Bose Corporation Joseph_Oparowski-at-bose.com
The Mountain, M/S 415 Phone: (508) 766-1371
Framingham, MA 01701-9168 Fax: (508) 766-1313


-----Original Message-----
} From: Michael L. Boucher [mailto:mboucher-at-tc.umn.edu]
Sent: Friday, June 07, 2002 1:52 PM
To: Microscopy-at-sparc5.microscopy.com


We have a Robinson Chamber View camera that we need to modify to fit a new
SEM. Does anyone know where I can contact the company or U.S. distributor?
Thanks for any replies.
Regards,
Mike
********************************************************************
Michael L. Boucher Sr. mboucher-at-tc.umn.edu
Lab Manager Rm 55 Office Ph 612-624-6590
I.T. Characterization Facility Rm 12 Main Off Ph 612-626-7594
University of MN Fax 612-625-5368
12 Shepherd Labs
100 Union Street S.E.
Minneapolis, MN 55455 http://resolution.umn.edu
********************************************************************








From daemon Mon Jun 10 07:41:32 2002



From: White, Woody N. :      nwwhite-at-mcdermott.com
Date: Mon, 10 Jun 2002 07:31:40 -0500
Subject: Glow Plug SEM analysis inquiry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Gary,

Allen had some good questions, but going ahead without all the answers...

Generally, the low-Z indications suggest carbon/hydrocarbon deposits which
likely occurred after the filament was at temperature. Did you do EDS? I
would suspect very high carbon, some oxygen, and a smattering of low
concentration elements. Ref: Glow 2-1, 2-10, Glow 3-12 This would not
have survived actual operational temperatures, I suspect.

Ref. Glow 3-1, 3-2, 3-3: Possible intergranular separation which is
(partially) filled with after-the-fact carbonaceous material. W wire is
typically exhibits long, fibrous like deformed grains.

In Glow 1-2, 1-3 I believe you are seeing grain channeling contrast. The
grains are more equiaxed here and I would suspect it was from an area that
was much hotter (not too hot) than the fibrous area.

In 1-4, the image suggests to me (in concert with the others) more low Z
material on the surface which did not play a role in the failure.


Woody White
McDermott Technology Inc.
McD: http://www.rdd.mcdermott.com/
Mine: http://woody.white.home.att.net


-----Original Message-----
} From: Gary Gaugler [mailto:gary-at-gaugler.com]
Sent: Friday, June 07, 2002 11:42 PM
To: MSA listserver


I am working on a rather fun subject of a model
engine glow plug. I have put some SEM pix at

http://www.microtechnics.com/glow.htm

which will pull up a PDF catalog of SEM images.
I think that the key images are the BSE ones,
including 10, 2, and 3. What I do not know is
what is going on in these pix. I tend to see a
phase change, but the metal coil wire is at room
temperature.

If anyone has any ideas about what is seen in
these pix, please reply off-line with your comments
and attribution data. As I am planning on writing
an article about this topic, I will include relevant
feedback and will attribution specifics.

One item of interest is that this particular plug was
relatively new (about 45 minutes of running from
totally new). It faltered and was replaced by a new
plug which solved the problem. so, why does this
happen and what exactly is happening to the plugs
to degrade them? Beats me.

gary g.



From daemon Mon Jun 10 07:41:53 2002



From: Susan Carbyn :      CarbynS-at-agr.gc.ca
Date: Mon, 10 Jun 2002 08:34:34 -0400
Subject: Edwards 306A Coating Instructions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi there,

If anyone has a set of instructions for downward evaporation of carbon using this machine, could I have a copy? We have three very different sets of instructions, none of which are very clear. We don't use liquid nitrogen with this machine.

Please contact me off line (or fax a copy) if you can help me out.

Thanks,

Susan



Susan Carbyn
Electron Microscopy Technician
Atlantic Food and Horticulture Research Centre
Agriculture and Agri-Food Canada
32 Main Street, Kentville, N.S. B4N 1J5
Canada

Phone (902) 679-5535
Fax (902) 679-2311

E-Mail: carbyns-at-em.agr.ca



From daemon Mon Jun 10 09:55:09 2002



From: Rosemary Walsh :      rw9-at-psu.edu
Date: Mon, 10 Jun 2002 11:38:58 -0500
Subject: Yur help re: Size of transferrin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,
Would any of you have a reference for the size of the
transferrin and the transferrin receptor?
Thanks for your help once again.
Rosemary

--
Rosemary Walsh, Manager
The Electron Microscope Facility for the Life Sciences,
A Shared Technology Facility, The Life Sciences Consortium
1 South FrearLab
Penn State University
University Park, PA 16802
(814) 865-0212
rw9-at-psu.edu
http://www.lsc.psu.edu/stf/em/home.html


From daemon Mon Jun 10 11:09:16 2002



From: S Keller :      swtkeller-at-yahoo.com
Date: Mon, 10 Jun 2002 09:01:16 -0700 (PDT)
Subject: TEM: Looking for a TEM w/ asid

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi:
I am looking for a TEM with STEM preferably a 200 Kv
JEOL.
Thx,
Sandra Keller

__________________________________________________
Do You Yahoo!?
Yahoo! - Official partner of 2002 FIFA World Cup
http://fifaworldcup.yahoo.com


From daemon Mon Jun 10 14:06:26 2002



From: Josh Kahn :      4jbk1-at-qlink.queensu.ca
Date: Mon, 10 Jun 2002 14:57:21 -0400
Subject: LM/TEM - cellular extraction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Can anyone recommend an extraction protocol to remove all cellular
material except intermediate filaments?

JBK
Read books




From daemon Mon Jun 10 15:49:26 2002



From: Don Grimes :      microtoday-at-mindspring.com
Date: Mon, 10 Jun 2002 14:56:34 -0500
Subject: Possible Virus

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Group,
Phil Oshel advises me that either my computer is infected with a virus, or
one of the people
I email is, and the virus is spoofing my address (pretending it was sent by
me/Ron Anderson, instead of the computer it really came from).
That address is microtoday-at-verizon.net and the subject line is: "Contar con
la carta de liberaci"
Not to be confused with my address, as above, or with that of Ron Anderson
(microtoday-at-attglobal.net), the new editor of Microscopy Today.
Regards to all,
Don Grimes



From daemon Mon Jun 10 17:14:54 2002



From: PHYSIOL4-at-AKAD.SUN.AC.ZA
Date: Mon, 10 Jun 2002 12:40:44 +0200
Subject: LM: need help fixing + embedding sugarcane

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi there

Does anyone have any experience with fixing and embedding
sugarcane stem tissue for in situ hybridization? So far we have
tried a number of protocols - including traditional fixation in 4%
paraformaldehyde, as well as prefixing and then freezing in
isopentane and cutting frozen sections - but to no avail. With
traditional chemical fixation (a protocol which worked beautifully for
one of our colleagues working on grape berries), we have to fix for 5-
7 days just to get adequate fixation, and even then parts of the
older internode tissue don't fix properly - this makes sectioning very
tricky. With cutting prefixed frozen tissue, our sections break up
very easily, and even when we do get sections, once thawed they
go into some sort of shock and develop thick black cell borders
(even in the xylem).

If anyone has any ideas, please let me know!

Thanks
Gabrielle



Gabrielle Turner
Institute for Plant Biotechnology
University of Stellenbosch
7600

Cell: 083 324 7453


From daemon Mon Jun 10 18:01:25 2002



From: David Hall :      hall-at-aecom.yu.edu
Date: Mon, 10 Jun 2002 18:54:35 -0400
Subject: lab microwave oven in Northwest Ohio

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This question is for a friend who is doing TEM in Toledo, Ohio. She
is hoping to get access to a lab nearby with an energy controlled
microwave oven (Pelco, for instance) where she could conduct
fixations from time to time.

You can direct responses to my email address. Many thanks.

Dave
--
David H. Hall, Ph.D.
Center for C. elegans Anatomy
Department of Neuroscience
Albert Einstein College of Medicine
1410 Pelham Parkway
Bronx, NY 10461

www.wormatlas.org
www.aecom.yu.edu/wormem

phone 718 430-2195
fax 718 430-8821


From daemon Mon Jun 10 18:16:59 2002



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 10 Jun 2002 16:10:21 -0700
Subject: RE: Glow Plug SEM analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I've received several very informative responses to my original
posting--many thanks to you all.

Some responses suggested that I am going through this
exercise for income purposes.....not. See:

http://www.photoweb.net/pw_trav/RC/RC.html

Nope....I'm doing this at a loss of money but a gain in fun and
education. Glow plugs are key items in 2- and 4-stroke
RC engines. Having a plug die in the middle of a competition
aerobatic maneuver is not at all good. Having one not start
on the ground when it's my turn is also not good. I'm newly
returning to RC as a hobby. Applying SEM and new
analytical technology to RC is very enjoyable.

It turns out that the glow plug wire is made of Platinum
and Rhodium. The Pt acts as a catalyst when introduced
with methanol. I don't know if there is a difference between
the 2- and 4-stroke plug wires. The veteran RC enthusiasts
are very knowledgeable about many aspects of this topic.

I put additional images at the site.

http://www.microtechnics.com/glow.htm


These added images are
of new 4-stroke glow plugs. They do not show the dramatic
grain boundaries seen in the failed 2-stroke plug. So, do
the boundaries show up after use, are they only seen in
2-stroke plugs, or what? Perhaps the 2-strokes are an
alloy whereas the 4-strokes are just Pt. More questions
and unknowns.....sigh.

But the difference in wire quality between the two 4-stroke
engine plugs is, I believe, significant. I have not used
these plugs yet. Now I will have to keep track of each
plug's experiences in order to make a valid experiment.
I had not thought about this much complication...

With Z of 78/Pt and 45/Rh, I ought to be able to see
the grain boundaries with BSE. I cannot do this with
a new plug. Therefore, as Warren and Woody have suggested,
there is the possibility of visualizing the effects of high
temperature exposure. I have seen this in ICs at Al/Si
alloy areas. But this was fundamentally when the
wafers were sintered too long or reflow was done
improperly.

One thing is clear however, there is marvelous depth of
field and clarity in looking down the coiled wire in the plug.
The SEM can do this with ease. LM has a very tough
time attempting this feat. The

This is becoming quite involved. Not sure how useful this is
other than to satisfy my own curiosity. There is perhaps
a way to electrically test new plugs, based on what I have
seen as clear defects in new plugs. However, considering
that a new plug costs about $7, it is probably not economical
to worry about pre-qualifying plugs! But making an assessment
of the quality of each brand and type does make sense.

I do wish that I had x-ray capability. I'm working on that.
There have been several good responses to my listserver posting
about that--thanks. I hope that an EDX system is not like
a boat--a place to dump lots of money. At least a dewar
is small and could be used as a boat anchor in desperation.

gary g.




At 05:31 AM 6/10/2002, you wrote:

} Hi Gary,
}
} Allen had some good questions, but going ahead without all the answers...
}
} Generally, the low-Z indications suggest carbon/hydrocarbon deposits which
} likely occurred after the filament was at temperature. Did you do EDS? I
} would suspect very high carbon, some oxygen, and a smattering of low
} concentration elements. Ref: Glow 2-1, 2-10, Glow 3-12 This would not
} have survived actual operational temperatures, I suspect.
}
} Ref. Glow 3-1, 3-2, 3-3: Possible intergranular separation which is
} (partially) filled with after-the-fact carbonaceous material. W wire is
} typically exhibits long, fibrous like deformed grains.
}
} In Glow 1-2, 1-3 I believe you are seeing grain channeling contrast. The
} grains are more equiaxed here and I would suspect it was from an area that
} was much hotter (not too hot) than the fibrous area.
}
} In 1-4, the image suggests to me (in concert with the others) more low Z
} material on the surface which did not play a role in the failure.
}
}
} Woody White
} McDermott Technology Inc.
} McD: http://www.rdd.mcdermott.com/
} Mine: http://woody.white.home.att.net
}
}
} -----Original Message-----
} } From: Gary Gaugler [mailto:gary-at-gaugler.com]
} Sent: Friday, June 07, 2002 11:42 PM
} To: MSA listserver
} Subject: Glow Plug SEM analysis inquiry
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Tue Jun 11 03:20:11 2002



From: Anaspec :      anaspec-at-icon.co.za
Date: Tue, 11 Jun 2002 10:05:15 +0200
Subject: ICEM 15 1 July registration date.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all

A quick update from South Africa and the organizing of ICEM15 for 1-6
September 2002.
The last chance to still save a bit on early registration is coming up fast.
1 July being the cut-off date. Registration is available online by visiting
the new revamped website. www.icem15.com

As the scientific programme has now been "finalised" there are some numbers
on the conference, available via the same website. Given all the events over
the past two years, the numbers of abstracts submitted is well over our
expectations.
The programme will be put onto the website once all last minute changes have
been made. In the mean time there is an indication of some of the topics of
interest.

The trade exhibition looks good as most stands are sold. As the trade have
been asking who and how many people will be there, we have added a page
showing the abstracts by country. Interesting reading!

Thanks for your time and hope to see you all at ICEM15 in Durban, South
Africa.

Luc Harmsen
Marketing ICEM15

www.icem15.com



ANASPEC South Africa www.anaspec.co.za
Tel: +27 11 794 8340
Fax: +27 11 794 8349
P.O. Box 2561
Honeydew 2040
Gauteng, South Africa




From daemon Tue Jun 11 08:36:39 2002



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Tue, 11 Jun 2002 09:22:33 -0400
Subject: Re: Yur help re: Size of transferrin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 11:38 AM -0500 6/10/02, Rosemary Walsh wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


From daemon Tue Jun 11 13:37:39 2002



From: Debby Sherman :      dsherman-at-purdue.edu
Date: 11 Jun 2002 13:27:58 -0500
Subject: Core Facility managment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Fellow microscopists:
This year we will discuss equipment maintenance during the Core Facility Management session at M&M2002. The session will be on Wed morning, Aug. 7 (although the program may say Wed afternoon - a mistake that hopefully was corrected). The program is at the end of this E-mail.

A number of the facilitators have asked for direction as to their comments. I would appreciate your sending specific questions or points of interest to me. Please indicate who you would like to address your issues. I will summarize the response and send the information on to the presenters. This should help make for a more productive session.

The NSF representative may not be able to make it to the meeting. However, if I have specific questions for NSF related to equipment purchase and their policies on funding of service contracts, I will try to get the information and present it at the session. Additional discussion and questions generated at the meeting will be clarified and sent out via the list.

We again hope to tape the discussion for later publication...providing I get the time to do the transcription.

See you in Quebec!
Debby

Maintaining Major Equipment in a Core Microscopy Facility

8:30: Servicing by Original Equipment manufacturers: the structure, function & considerations for assembling & operating a major service organization
JEOL: Patrick McGinley
National Service Manager
FEI: Mike Kearney
Director of Service
Hitachi: Greg Rigby
Director of Service

9:15: Servicing EMs by On-site staff: Insight into the Do-it-yourself Approach.
John Wheatley, Arizona State University
Owen Mills, Michigan Technological University

10:00: Break

10:30: Using 3rd Party Service Organizations for Major Equipment Maintenance
Art McCanna
Service Manager
Materials Analytical Services

11:00: Policies Regarding Allowable Costs for Equipment Maintenance on National Science Foundation (NSF) Awards

Debby Sherman Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
West Lafayette, IN 47907





From daemon Wed Jun 12 07:44:24 2002



From: Lesley S. Bechtold :      lsb-at-jax.org
Date: Wed, 12 Jun 2002 08:27:51 -0400
Subject: Salary Survey for Microscopists

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id HAA11853
for dist-Microscopy; Wed, 12 Jun 2002 07:33:23 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id HAA11849
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Wed, 12 Jun 2002 07:32:52 -0500 (CDT)
Received: from aretha.jax.org ([192.43.249.2])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id HAA11842
for {microscopy-at-sparc5.microscopy.com} ; Wed, 12 Jun 2002 07:32:36 -0500 (CDT)
Received: from spikey.jax.org (spikey.jax.org [199.94.155.39])
by aretha.jax.org (8.9.1/8.9.1) with ESMTP id IAA24237
for {microscopy-at-sparc5.microscopy.com} ; Wed, 12 Jun 2002 08:27:52 -0400 (EDT)
Message-Id: {5.1.0.14.0.20020612082621.00a73dd0-at-aretha.jax.org}
X-Sender: lsb-at-aretha.jax.org
X-Mailer: QUALCOMM Windows Eudora Version 5.1


Two or three years ago, "Microscopy Today" conducted a salary survey for
microscopists based on years of experience, education and region. Does
anyone know if that was archived anywhere? Thank you!

Lesley Bechtold



Lesley S. Bechtold
Supervisor, Biological Imaging
The Jackson Laboratory
600 Main St.
Bar Harbor, ME 04609
207-288-6191



From daemon Wed Jun 12 09:31:38 2002



From: Gillmeister, Russ :      RGillmeister-at-crt.xerox.com
Date: Wed, 12 Jun 2002 10:23:15 -0400
Subject: Roughness from IPP scan

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings Microscopists:

I would like to measure roughness from a microtomed cross section of a
surface replica. We're using Image Pro Plus to trace the interface
generating a representation of the surface profile. Is there some freeware /
software available to get from the trace generated by IPP to some
quantification like Ra???

Thanks,

Russ Gillmeister
Microscopy
Xerox Corp.
Webster, NY 14580




From daemon Wed Jun 12 10:51:27 2002



From: woxberry-at-downstate.edu
Date: Tue, 11 Jun 2002 22:21:15 -0500
Subject: question on cryoultramicrotomy equipment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Listers,

I am trying to get an old FC-4 cryo unit up and running to do some
cryosectioning on my Ultracut E. It seems to be missing some parts,
specifically the LN2 transfer hose from the dewar to the unit and a male
threaded transfer tool to transfer the specimen holder into the
chamber. I only
have one knife holder .

Does anyone have an FC-4 gathering dust somewhere who might want to find it a
new home? Or at least the LN2 transfer hose? I found one source for
a hose but
they want over $600 ! I don't want to buy a hose and find out that the unit
doesn't work.
Thank you all.

Bill Oxberry
Core Microscopy Lab
Dept. of Path.
Downstate Medical Center
Brooklyn, NY
woxberry-at-downstate.edu
7182704472


From daemon Wed Jun 12 13:40:12 2002



From: Thearith H. Ung :      tung-at-qdots.com
Date: Wed, 12 Jun 2002 11:28:18 -0700
Subject: RE: Urylacetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Would anyone be interested in having a solution of urylacetate (25g)? We
have a bottle that has not been opened since it was purchased. We would be
happy to ship it out to anyone interested. Please let me know as soon as
possible if you would to have it.

Regards,
Thearith
_________________

Thearith Ung

Quantum Dot Corporation
26118 Research Road
Hayward, CA 94545, USA
Tel: 510-887-8775 (Ext 4125)
Fax: 510-783-9729
Email: tung-at-qdots.com



From daemon Wed Jun 12 13:58:41 2002



From: Eric Anderson :      anderson_e-at-southernct.edu
Date: Wed, 12 Jun 2002 14:52:29 -0400
Subject: Re: Philips EM400 Matching Mains Transformer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings Vitaly,

Sorry to trouble you again, but I think I may have goofed. When I told our
campus electricians to install a 220V service for the EM400 (before I had the
manuals), they naturally gave us two hots and a neutral. From each hot to
neutral is about 118V. From hot to hot is only 205V. When I asked the
microscopy group for help with my 237V, I was merely adding the two separate
measurements with respect to neutral, and worse, I had not considered that
Philips asks for a 3-wire hook up consisting of 220V/neutral/ground, not
110/110/neutral.

Since the EM400 has a power spec of 8 kVA and the Buck & Boost only 1 kVA, I
called the transformer manufacturer Square D, and they say I need a custom
built $2000 isolation transformer to get the mains set up properly (without the
Buck&Boost), or else a $1200 stock gizmo to use with the Buck&Boost. I already
have the Buck & Boost, but sadly, our budget is virtually exhausted. Our
electrician thinks we should try using the two hots in place of the
220/neutral, but I'm very reluctant. Do you operate an EM400, and how is yours
powered?

Thanks so much,
-Eric
---------------------------
Eric Anderson
SCSU Physics Adjunct
203-392-6455
anderson_e-at-southernct.edu
---------------------------
Vitaly Feingold wrote:

} Eric,
}
} 237V is a bit too high. It must be between 208V and 220V. The simplest and
} economical way to power up your TEM is to use Buck and Boost transformer
} connected as autotransformer. Philips matching unit originally supplied with
} EM400, was also an autotransformer. Order from Grainger, www.grainger.com ,
} many locations everywhere, stock # 1H270, $174. This one will reduce your
} line voltage by 24V, making it 213V to 211V (see below), which is perfect.
}
} Notes.
} 1) Make sure that grounding is correct and safe.
} 2)This unit can be connected in 4 different ways, for increasing or reducing
} the line voltage by 12V or 24V, and is shipped with 1 page manual. Read it.
} 3) The above part number is given for your particular (line voltage) case.
} Others may require a transformer with different part number.
} 4) Transformer will not stabilize the line voltage, only change it.
} 5) I assume that you measured line voltage with no load connected- the
} actual voltage may drop about 1V or 2V when you connect the TEM.
}
} Vitaly Feingold
} Scientific Instruments and Applications
} 2773 Heath Lane, Duluth GA 30096
} (770)232-7785 ph.
} (770)232-1791 fax
} (678)467-0012 mobile
}
} This message is made of 100% recycled electrons.
} ----- Original Message -----
} From: Eric Anderson {anderson_e-at-southernct.edu}
} To: {Microscopy-at-sparc5.microscopy.com}
} Sent: Tuesday, June 04, 2002 12:53 PM
} Subject: Re: Philips EM400 Matching Mains Transformer
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } } Greetings All!
} } }
} } } We are in the process of setting up a second-hand EM400 acquired
} } } recently, and have found that the mains matching transformer was not
} } } included. Our raw supply main is 237 volts/60Hz (118V on one leg,
} } } 119V on the other, stability unknown), and I'm thinking this is not
} } } close enough to the specified 220V to go without the transformer. Any
} } } ideas? If we do need some line conditioning, can anyone recommend a
} } } particular device, or source for the original Philips transformer?
} } }
} } } Many thanks for any tips!
} } } -Eric



From daemon Wed Jun 12 14:43:26 2002



From: Pierre-M. Charest :      pcharest-at-rsvs.ulaval.ca
Date: Wed, 12 Jun 2002 15:32:30 -0400
Subject: M&M 2002 - Quebec City

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Colleagues Microscopists

The count down has started. In less than two months, Microscopy &
Microanalysis 2002 will begin in Quebec City, 4-8 August 2002. The
deadline for pre-registration is July 6. You can register on-line at
the following URL address:

http://www.microscopy.com/MSAMeetings/MMMeeting.html

More that 800 papers and posters will be presented at this meeting.
The scientific and social program will certainly please you. This
meeting could also be a unique opportunity to visit the French cradle
of North America. On August 7 the "Festival de la Nouvelle France"
will be launched, you will be witness of the life style of the early
French and English citizens who colonized our land from the XVIIth
Century. The old city will be crowded with people dressed like three
centuries ago.

If you visit Quebec with your family or friends, do not miss the
Whale Watching Cruise organized either on Wednesday August 7 or on
Friday August 9. You can make reservation on-line on the web site of
the Local Arrangement Committee where several tips are also offered
for your travel to Quebec, your lodging and your sightseeing
activities.

http://msc.rsvs.ulaval.ca/2002/2002.html

All members of the Local Arrangement Committee are eager to
facilitate your stay in our home and we hope we'll have the pleasure
to seeing you here.

AU REVOIR


PIERRE, CHAIR
LOCAL ARRANGEMENT COMMITTEE



From daemon Wed Jun 12 15:02:31 2002



From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Wed, 12 Jun 2002 14:56:02 -0500
Subject: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,

We have an Emitech cryopreparation unit associated with our Hitachi S4700 FESEM. The cryo unit's sputter coater is currently outfitted with a gold target, but we'd like to try platinum for a finer grain coating. The current target appears to be a thick sheet of gold, wrapped around the support and glued in place.

My main question: is there any reason we couldn't purchase a similar, malleable sheet of platinum and just wrap it around the gold for when we want to coat with Pt? Seems to me like it would work, but I like to check these things with the experts first.

Also, if anybody has any wisdom to share on avoiding getting ice on samples during cryo runs, I'd love to hear it. We often (usually) get ice during transfers and need to sublime off the moisture before coating, otherwise the coating forms quasi-permanent ice "casts" that render the sample unusable. We've replaced/checked all the o-rings we can reach, so any special hints are very welcome.

Finally, has anyone ever tried mixing colloidal carbon (like SEM mounting carbon) with the polyvinyl alcohol cryo mounting medium in order to improve conductivity? We're going to give it a whirl, but it would be interesting to hear if others have tried it.

Thanks!

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/




From daemon Wed Jun 12 20:10:21 2002



From: Rosemary Walsh :      rw9-at-psu.edu
Date: Wed, 12 Jun 2002 21:04:43 -0700
Subject: Re: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} HI Randy,

I use an Oxford (now GATAN) C1500C system and have some experience
imaging and analyzing cells grown in inserts, tissues, polymers,
emulsions and foods. When possible I prefer unfixed tissue or fixed,
cryoprotected cells.

} Also, if anybody has any wisdom to share on avoiding getting ice on
} samples during cryo runs, I'd love to hear it. We often (usually) get ice
} during transfers and need to sublime off the moisture before coating,
} otherwise the coating forms quasi-permanent ice "casts" that render the
} sample unusable. We've replaced/checked all the o-rings we can reach, so
} any special hints are very welcome.



In order to insure some success in removing surface ice, I blot excess
fluid from a 2mm sample, plunge into LN in a transport dewar located as
close to the SEM sample prep chamber as it possible, cover and quickly
press the N admit to open the chamber. After inserting the transfer holder
onto the prep cryostage (-170C), I pull a vaccum, wait 2 min., fracture if
needed, transfer onto the SEM coldstage, close the ball valve and
immediately begin heating the stage.I prefer to do the transfer this way
rather than into the LN slush where I run the risk of heavy ice
contamination. Etching or sublimation can take 15 min for plant tissues,
biopolymers to 45 minutes for emulsions and frozen foam foods (yogurt, ice
cream).

If your sample size is large 5-10 mm, I attach to the sample holder with
graphite/OCT drop and immediately insert onto sample prep stage (-170C) and
proceed as outlined above---I've also modified this by progressively
lowering temp using ice, dry-ice, LN plunge.
One adaptation for dry-ice use is to cut out a hole in the side of a
styrofoam container so that the transfer rod can be inserted into dry ice
vapor. It is the only way I to work with frozen samples.

It's is important to minimize the sample size--I use gold planchets from an
old Balzers FF/FE system -- they have a tiny depression which holds 10
uls. This sits in a brass holder with an aluminum cap which holds the
planchet in place. I invert a second and transfer to the cryo-prep stage,
wait two minutes, fracture by touching the pick to the top planchet and
proceeding as above.
Another option is to use brass rivets--the machine shop made 10 mm brass
stubs with holes drilled for three rivets--I either drop liquid emulsion or
scoop ice cream into them, cover with an inverted- cooled rivet, insert
onto the cryo prep stage, pull a vacuum, tip the top rivet so that it
fractures, insert into the SEM cryo-stage to begin sublimation and proceed
as above.

It is imperative for us to schedule cryo-SEM work during lower humidity
times. July and August are difficult times. If I have to work then, I run
a dehumidifier but it is still problematic.


} Finally, has anyone ever tried mixing colloidal carbon (like SEM mounting
} carbon) with the polyvinyl alcohol cryo mounting medium in order to
} improve conductivity? We're going to give it a whirl, but it would be
} interesting to hear if others have tried it.

I routinely mix a fresh batch of OCT cryo-mountant with a couple of drops
of colloidal graphite, place a small drop onto the brass mount, add a tiny
wedge of filter paper to the drop along with the sample. I use the filter
paper as a monitor for the sublimation / etch time--it has worked very well
with pieces of tissue, polymers. It is important to make this fresh
conductive adhesive, keep it in a one ml microfuge tube and apply with a
wooden pick or with a disposable 1 cc syringe.

I wish you success----ice can be maddening
Rosemary

Rosemary Walsh, Manager
The Electron Microscope Facility for the Life Sciences,
A Shared Technology Facility, The Life Sciences Consortium
1 South FrearLab
Penn State University
University Park, PA 16802
(814) 865-0212



From daemon Thu Jun 13 01:46:36 2002



From: DrJohnRuss-at-aol.com
Date: Thu, 13 Jun 2002 02:34:49 -0400
Subject: Re: Roughness from IPP scan

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In a message dated Wed, 12 Jun 2002 10:43:57 AM Eastern Daylight Time, RGillmeister-at-crt.xerox.com writes:

} I would like to measure roughness from a microtomed cross section of a
} surface replica. We're using Image Pro Plus to trace the interface
} generating a representation of the surface profile. Is there some freeware /
} software available to get from the trace generated by IPP
} to some
} quantification like Ra???

The Fovea Pro (ReindeerGraphics.com) plugins run in Image Pro Plus version 4.5 and include this function. Besides Ra you also get Rq and some other standard profile roughness values.





From daemon Thu Jun 13 03:14:05 2002



From: Chris Jeffree :      c.jeffree-at-ed.ac.uk
Date: Thu, 13 Jun 2002 09:06:57 +0100
Subject: Re: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Randy
It is possible to make up your own targets. If you buy discs of
precious metal sheet from a reputable dealer you will usually save a
considerable amount of money compared with the sputter manufacturer's
prices. And you can specify target thickness. But make sure the metal
purity is very high.
Although you could wrap platinum round the gold for a temporary fix,
it would probably be a better plan to buy a spare target base and make
that up with your platinum sheet. It is important to ensure secure
electrical connection between the base and the metal foil. The easiest
way to do this is to glue the foil to the base with silver-loaded
epoxy. SPI, TAAB, Agar etc. can usually supply this.

Penny- (or dime-) pinchers tip: Sputter coaters often burn through
their targets in a narrow ring leaving a huge amount of perfectly
good metal in the centre and round the edges. When this happens with
your second target you can use the gold saved from the first to patch
the holes.

} We have an Emitech cryopreparation unit associated with our Hitachi
S4700 FESEM. The cryo unit's sputter coater is currently outfitted
with a gold target, but we'd like to try platinum for a finer grain
coating. The current target appears to be a thick sheet of gold,
wrapped around the support and glued in place.
}
} My main question: is there any reason we couldn't purchase a
similar, malleable sheet of platinum and just wrap it around the gold
for when we want to coat with Pt? Seems to me like it would work, but
I like to check these things with the experts first.

It is virtually impossible to avoid some contamination of exposed
surfaces with ice during cryofixation and the initial transfer stage.
The first source is particulate ice in the nitrogen slush, condensed
from the atmosphere when the nitrogen is poured into the dewar.
Usually this ice is in the form of flocculent particles that are
easily sublimed at -100oC or above. Ice condensed on the cold specimen
directly from the moist atmosphere will also be finely granular, and
easily sublimed. This type of ice is what you will normally observe if
there is a leak in your transfer system seals. If you are seeing
encasing films or sheets of ice covering the specimen there was
probably a film of liquid water covering the specimen before it was
cryo-fixed. This is a major problem when examining the surfaces of
cultured cells, for example. The only way I can think of to deal with
this if you want to observe fully or partially hydrated material is to
blot the surplus water off thoroughly immediately before fixation, and
then etch some ice off. The problem with this is that the etch times
will be quite long, and since the film will be of very uneven
thickness very variable amounts of freeze-drying will occur in the
specimen.

} Also, if anybody has any wisdom to share on avoiding getting ice on
samples during cryo runs, I'd love to hear it. We often (usually) get
ice during transfers and need to sublime off the moisture before
coating, otherwise the coating forms quasi-permanent ice "casts" that
render the sample unusable. We've replaced/checked all the o-rings we
can reach, so any special hints are very welcome.

Aqueous carbon dag is quite effective as a mountant on its own or when
mixed with Tissue-Tek. Dirty though! I have used a 50:50 mixture, but
the effective proportions will depend on the solids content of your
carbon dag preparation. You could try measuring the resistivity of the
frozen mixtures. I don't recommend using SEM mounting carbon dag of
the Leit C type which uses an organic solvent.

Chris
} Finally, has anyone ever tried mixing colloidal carbon (like SEM
mounting carbon) with the polyvinyl alcohol cryo mounting medium in
order to improve conductivity? We're going to give it a whirl, but it
would be interesting to hear if others have tried it.
}
} Thanks!
}
} Randy
}
} Randy Tindall
} EM Specialist
} Electron Microscopy Core Facility
} W122 Veterinary Medicine
} University of Missouri
} Columbia, MO 65211
} Tel: (573) 882-8304
} Fax: (573) 884-5414
} Email: tindallr-at-missouri.edu
} Web: http://www.biotech.missouri.edu/emc/
}
}
}




From daemon Thu Jun 13 08:29:22 2002



From: TCLEEG-at-tsmc.com.tw (by way of MicroscopyListserver)
Date: Thu, 13 Jun 2002 08:10:49 -0500
Subject: Senior TEM analyst position at tsmc

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



}
}
}
} Senior TEM analyst
}
} Taiwan Semiconductor Manufacturing Company (TSMC) is seeking
} qualified candidates for the position of Senior Failure Analysis
} Engineer for the Process Failure Analysis Department located in
} Hsin-Chu, Taiwan. The position involves TEM analysis of
} reliability failures and customer returns using tool such as
} JEOL2010F, JEOL 2000 EX and FIB DB835 etc. The successful
applicant
} will also be responsible for developing solutions to technical
} problems and issues in the failure analysis group, including the
} analysis and interpretation of results to resolve
} product/process and customer problems, independent investigative
} work, new
} analytical technique development and presentation of results in
} customer and technical meetings.
}
} Applicant Qualifications:
} - High level of knowledge and hands-on experience in TEM analysis.
} - Ability to conduct independent laboratory investigations.
} - Good oral and written communication skills as required in
} preparing reports on TEM results and problem investigation.
} - Experience with semiconductor materials analysis and sample
} preparation is a plus.
}
} Interested individuals should send a detailed cover letter and
} resume to: Tan-Chen Lee, Process Failure Analysis Department,
} Taiwan Semiconductor Manufacturing Co., 25 Li-Hsin Rd.
} Science-Based Industrial Park Hsin-Chu, Taiwan 300 or via
} e-mail:tcleeg-at-tsmc.com.tw
}
}
}
} Best regards,
}
} Tan-Chen Lee ???
} Process Failure Analysis Department
} Taiwan Semiconductor Manufacturing Company, Ltd.
} 25, Li-Hsin Rd, Science-Based Industrial Park
} Hsin-Chu, Taiwan, R.O.C.
} tcleeg-at-tsmc.com.tw
}
}


From daemon Thu Jun 13 08:50:36 2002



From: Debby Sherman :      dsherman-at-purdue.edu
Date: 13 Jun 2002 08:44:12 -0500
Subject: Re: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
"message to: MSA list" {microscopy-at-sparc5.microscopy.com}
X-Mailer: QuickMail Pro 2.1 (Mac)
X-Priority: 3
MIME-Version: 1.0
Reply-To: Debby Sherman {dsherman-at-purdue.edu}
Content-Type: text/plain; charset="US-Ascii"
Content-Transfer-Encoding: 8bit
X-MIME-Autoconverted: from quoted-printable to 8bit by sparc5.microscopy.com id IAA16742



From: Debby Sherman :      dsherman-at-purdue.edu
Date: 13 Jun 2002 08:44:12 -0500
Subject: Re: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Randy,
We have mixed carbon with Tissue Tec OCT cryoadhesive since the early 80's for cryo-SEM. Formulation we use is:
0.4g carbon powder (can be obtained from waste from sharpening carbon rods used for carbon coating)
2gm 95% ETOH
10gm OCT compound
Mix well and let sit for a few days prior to using.

Also, I use an 18 year old Hexland/Oxford/Gatan system for plunge freezing/coating etc. It is lower tech than today's versions but still does imazingly well. We have minimum problem with ice on samples accumulated during transfer. The only time I sublimate is if I have to fracture a sample and remove some of the water from the fracture surface to better reveal the structure. Now this is with a standard tungsten-filament instrument so maybe we are not seeing the water. However, it is not visible at magnifications up to ~5000x, which is about as high as we can go with biological samples before excessive noise swamps the signal and we get into the "empty magnification" problems.
I am not familiar with the EMItech accessory so cannot comment on possible reasons or solutions for this ice problem but assume the company may have some suggestions. Also, perhaps they can manufacture a target for you from another metal. However, wouldn't it require different voltage and possibly vacuum conditions to sputter platinum instead of gold? I would appreciate hearing if you manage to do this as it will be of value to many of use who do cryo and hope to do it on FESEM's in the future.

Debby


Debby Sherman Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
West Lafayette, IN 47907


On Wednesday, June 12, 2002 2:56 PM, Tindall, Randy D. {TindallR-at-missouri.edu} wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Thu Jun 13 09:15:14 2002



From: Microscopy Today :      microtoday-at-attglobal.net
Date: Thu, 13 Jun 2002 10:08:52 -0400
Subject: Salary Survey for Microscopists

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id JAA16901
for dist-Microscopy; Thu, 13 Jun 2002 09:14:11 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id JAA16898
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Thu, 13 Jun 2002 09:13:41 -0500 (CDT)
Received: from prserv.net (out2.prserv.net [32.97.166.32])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id JAA16891
for {microscopy-at-msa.microscopy.com} ; Thu, 13 Jun 2002 09:13:29 -0500 (CDT)
Received: from ibmav8jk6vcief (slip-12-64-152-46.mis.prserv.net[12.64.152.46])
by prserv.net (out2) with SMTP
id {2002061314085020206qbc4me} ; Thu, 13 Jun 2002 14:08:50 +0000
Reply-To: {microtoday-at-attglobal.net}


The Microscopy Today archives for the last five years are on the MT
website, www.microscopy-today.com
I have copies of most of the old issues.

Ron Anderson, MT Editor

-----Original Message-----
} From: Lesley S. Bechtold [mailto:lsb-at-jax.org]
Sent: Wednesday, June 12, 2002 8:28 AM
To: microscopy-at-sparc5.microscopy.com


Two or three years ago, "Microscopy Today" conducted a salary survey for

microscopists based on years of experience, education and region. Does
anyone know if that was archived anywhere? Thank you!

Lesley Bechtold



Lesley S. Bechtold
Supervisor, Biological Imaging
The Jackson Laboratory
600 Main St.
Bar Harbor, ME 04609
207-288-6191





From daemon Thu Jun 13 12:14:03 2002



From: eld26-at-cornell.edu
Date: Thu, 13 Jun 2002 13:04:13 -0400 (EDT)
Subject: Polishing trabecular bone for AFM/nanoindentation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Listers,

I have some samples of cancellous bone embedded in both spurrs and PMMA,
and I plan to make small indentations (1-2 microns) within a single
lamella. To do this, though, I need a relatively smooth (~ {10 nm rms
roughness on a 20 um scan) and uniform surface.

My current surface prep procedure of manually grinding on SiC paper
through 1200 grit and then polishing with Al2O3 on hard nylon cloth down
to 0.05 um particle size results in samples that are generally too rough
(~70-120 nm rms roughness on a 20 um scan), and the surface finish is
highly variable from point to point within a sample.

Does anyone have suggestions for improving the surface finish?
Unfortunately, I don't have access to automatic polishing equipment, and
I'm wary of chemical treatments that may alter the surface properties I'm
trying to measure.

thanks very much,


Eve Donnelly
Sibley School of Mechanical & Aerospace Engineering
Cornell University
Ithaca, NY 14850
tel: (607)255-3582




From daemon Thu Jun 13 14:12:37 2002



From: Lesley S. Bechtold :      lsb-at-jax.org
Date: Thu, 13 Jun 2002 15:04:32 -0400
Subject: Uranyl Formate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,

I'm posting this for a microscopist not on the listserver. They
are testing uranyl formate as a negative stain without much success. Does
anyone have any tips or methods for getting this to work? Thanks!

Lesley Bechtold

Lesley S. Bechtold
Supervisor, Biological Imaging
The Jackson Laboratory
600 Main St.
Bar Harbor, ME 04609
207-288-6191



From daemon Thu Jun 13 15:34:47 2002



From: hseyan :      hseyan-at-is.dal.ca
Date: Thu, 13 Jun 2002 17:26:18 -0300
Subject: Double immuno staining for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
I would like to know if there is anybody who has successfully managed to
immunostain nerve cells in brain tissue labeled with Cholera toxin (Ctb) using
antibodies against CTB,developing with TMB method, and then follow it with
antibody against GABA using a goldconjugated secondary. The preparation is to
be viewed using a TEM. Please advise.

harjit Seyan
hseyan-at-is.dal.ca




From daemon Thu Jun 13 17:06:53 2002



From: Lewis Melissa A :      LEWISMA-at-medicine.ufl.edu (by way of
Date: Thu, 13 Jun 2002 16:55:44 -0500
Subject: glass knives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am having problems with my water level not staying on the knife
edge. I am cutting Lowicryl, so the water level has to be very low.
This problem has occurred with the last two batches (10-15) of knives
that I have made. I have tried everything I know with no success.
Does anyone have any suggestions. Please:)


From daemon Thu Jun 13 17:06:58 2002



From: Robert Johnson :      spore1-at-worldnet.att.net (by way of
Date: Thu, 13 Jun 2002 16:56:16 -0500
Subject: Greetings EM in D.C.?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friends,

I am an early career Biologist and photographer who is new to the
Washington, D.C. area and interested very much in working in the field of
electron microscopy and wish to make myself available to labs needed extra
help. Should anyone be able to provide advice or contact information of
parties willing to meet with me and tour their facility, I would be very
grateful.


-Robert Johnson

(202) 265-2264

spore1-at-worldnet.att.net


From daemon Thu Jun 13 18:18:07 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Thu, 13 Jun 2002 16:11:01 -0700
Subject: Re: Uranyl Formate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Lesley

Use uranyl formate fresh (1-2 h after preparation) and avoid exposure to
the direct light (alumina foil wrap when dissolved, diffuse room light when
prepare samples - no direct high intensity light). I am using 1% UF
dissolved in ultrapure H2O for 30-40 min with gentle mixing in the
dark. Staining time 1-2 min. The quality of staining depends from the
support film (I am using 1.5 nm thick carbon film), sample
purity/concentration. In most cases I got staining quality better that
with UA. UF in general produces fine granularity and looks less contrast
under the microscope, but it went fine on the film.

I hope it help. Sergey

At 12:04 PM 6/13/02, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Thu Jun 13 18:51:37 2002



From: Rosemary White :      rosemary.white-at-csiro.au
Date: Fri, 14 Jun 2002 09:48:23 +1000
Subject: RE: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Randy,

Yes the slow-freeze process is very simple, exactly as you say - we just
attach the plant material to any type of stub with carbon dag and freeze on
the cold preparation stage in the coating/prep chamber. This almost
completely eliminates any frost, in fact we rarely check for frost. If
it's a very precious specimen we check the surface at 2-3 kV before
coating. The only problem we have is if the plant tissue is broken and
exudes cell contents before freezing. If this is going to be a problem we
just work very quickly and mop up gently - for example, if we want to look
at the innards of a flower after removing some of the sepals and petals, or
if we want to look at a stem or root cross-section. We also do Arabidopsis
siliques like this to look at ovule size and morphology in various mutants
- slit in half, glue down halves onto stub, freeze, coat, observe. The
systematists/taxonomists in the Herbarium also look at all their fresh
material this way. Another group working on the soil-plant interface look
at roots in soil in this way too. Yes, we put dirt in our SEM!

This is a joint facility with the Entomology Division of CSIRO, and they do
the same for their fresh insects. With small insects, they knock them out
with CO2 I think, line up lots of them on a large stub, freeze on cold
stage of prep chamber, coat, and spend the rest of the day taking photos.

One advantage we have is that Canberra is quite dry, rarely gets above 80%
humidity in winter. Summer can be down to 20%.

With already frozen tissue then we routinely sublime off the frost by
raising the stage temperature to -90 while observing the tissue at low kV.
We drop the temperature just before all the frost has gone - have to watch
quite closely not to etch cryo-planed specimens too much. This is only for
tissue that has been frozen in the field and sent in for analysis, or for
tissue that has been cryo-planed for X-ray analysis. We attach this with
TissueTek or Leit-C. The tissue for cryo-planing is inserted into
home-made chucks that go into the cryomicrotome (home-made - by our machine
shop - is much cheaper!). Then for observation two chucks are inserted
into a specially made stub to fit onto the cryostage.

A final comment re Pt or other coating, for our Oxford 1500 CT cryochamber
we have two spare target bases (cf. comment from Chris Jeffree), one for
carbon rope, the other for W wire. As I mentioned before, for Al coating
we wrap pure Al sheet around the W wire (40 amp double strand). Al is from
Alfa Aeser, who have a vast range of pure metal bits and pieces (no
commercial interest........). You have to find the right Ar level, etc.
for good Al coating - by trial and error. But once this is sorted out,
it's quite routine. The expert on cryoplaning and Al (and Cr) coating for
X-ray analysis in Canberra is Cheng Huang, who has published quite a few
papers using these methods.

We get all our gold targets made by a local jeweller - we get them made
thicker than normal (0.6mm), and we take back the bits from worn out
targets so he can remelt this and incorporate into the new target. We've
found this to be much cheaper than getting the gold from any other source
and it only takes a day or so. I guess because any other gold targets have
to be imported, it makes a big difference here in Oz. And the jeweller has
some gold-coated insects on display in his shop....

hope this is of some use!
cheers,
Rosemary

} Hi Rosemary,
}
} Thanks for the information! I'm a little unclear about your slow-freeze
} process, though. Could you give me a little more detail on how you freeze
} the sample in the transfer chamber. On our unit, there is a freezing
} chamber for the LN2 cup, and the transfer chamber is just a holder that
} can be pumped to a vacuum to move the sample from the freezing chamber, to
} the coating/preparation chamber and then to the SEM. Do you freeze the
} sample by simply setting on the LN2-cooled preparation stage?
}
} Thanks again.
}
} Randy
}
} -----Original Message-----
} From: Rosemary White [mailto:rosemary.white-at-csiro.au]
} Sent: Thursday, June 13, 2002 12:46 AM
} To: Tindall, Randy D.
} Subject: Re: SEM/ Cryo unit sputter coater
}
}
} Hi Randy,
}
} Can't comment on Pt, but for Al coating we do exactly that - wrap a very
} thin sheet of Al around a W electrode for sputtering frozen plant material
} before X-ray analysis.
}
} Re. icing up - I guess this is transfer from freezer to cryochamber(?),
} have you ever tried "slow-freezing"? We do this all the time for plant
} material - just stick the material onto the cryo-stub with Tissuetek or
} that liquid carbon goop or whatever, then put into transfer chamber, and
} freeze. Then check for ice, usually the specimen is fine, then withdraw to
} chamber again, coat, and observe. Works for robust things, but I guess
} would not work for some animal tissues.
}
} cheers,
} Rosemary
}
} }
} }


Rosemary White
Microscopy Centre
CSIRO Plant Industry
GPO Box 1600
Canberra, ACT 2601
Australia

ph. 61- 2 6246 5475 or
mob. 61- 0402 835 973
rosemary.white-at-csiro.au




From daemon Thu Jun 13 22:55:29 2002



From: Sampath, Srinidhi (CORP, GEITC) :      Srinidhi.Sampath-at-geind.ge.com
Date: Fri, 14 Jun 2002 09:17:02 +0530
Subject: Characterizing inhomogeneous particle distributions...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for techniques, protocols and programs which will give a
quantitative measure of particle distribution in a matrix. Emphasis is on
inhomogeneous distributions and how to characterize them.
I would appreciate any information on this. You could also reply to my
e-mail address.

Thanks,

S. Srinidhi,
Materials Scientist
John F. Welch Technology Center
_______________________________________
Materials Research Laboratory
Sy #152, Export Promotion Industrial Park Phase - 2
Hoodi Village, Whitefield Road,
Bangalore - 560066.
Phone: +91-80-8412050 - 69 x: 2562
Fax: +91-80-8412111
e-mail: srinidhi.sampath-at-geind.ge.com
Dialcom: 8*901 2562


"THIS E-MAIL MESSAGE ALONG WITH ANY ATTACHMENTS IS INTENDED ONLY FOR THE
ADDRESSEE and may contain confidential and privileged information.
If the reader of this message is not the intended recipient,
you are notified that any dissemination, distribution or copy of this
communication is strictly Prohibited.
If you have received this message by error, please notify us
immediately, return the original mail to the sender and delete the
message from your system."



From daemon Thu Jun 13 23:29:22 2002



From: jerzy.gazda-at-amd.com
Date: Thu, 13 Jun 2002 23:22:48 -0500
Subject: Bacteria on semiconductor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Experts,

as a materials scientist I don't have too much exposure to bio-elemental analysis. Could someone in Bio-fields help out. What elements should I expect to find in biological residue? I am evaluating suspected bacterial deposits (contamination) and finding O, C, and Ca. Would there be any other elements present after the residue was encapsulated in high temperature deposited silicon oxide?

Thank you in advance.

Warmest Regards from Texas.

Jerzy



******************************************************
Jerzy Gazda, Ph.D. Advanced Micro Devices
Supervising Engineer 5204 E. Ben White Blvd. - MS 613
PCAL - AIM Section Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453 FAX: (512) 602-7470
jerzy.gazda-at-amd.com
******************************************************





From daemon Fri Jun 14 02:54:01 2002



From: heiko.stegmann-at-amd.com
Date: Fri, 14 Jun 2002 09:48:19 +0200
Subject: Re: Bacteria on semiconductor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rosemary
While I concede the slow-freeze method works on some specimens - it
can as you indicate be
particularly successful with leaves or small insects that are covered
with a cuticle -
such recidivism should only be practised by consenting adults fully
aware of the implications, which are
a) wet stuff (possibly including organic solvents) has to be
introduced to the high vacuum of the prep chamber thus contaminating
cold stage and anticontaminator of the prep chamber
b) uncontrolled evaporation from the surface of the un-fixed specimen
c) thus specimens without the benefit of a protective covering (such
as cultured cells) will show variable amounts of drying artefacts.
I suppose the frank reality is that all etching / drying procedures
result in artefacts, but there is a comforting illusion of control if
you can get the same artefacts with reasonable reproducibility!

On the topic of gold for targets, I was not recommending jewellers as
a source of metals (although that is fine if they, like yours, seem to
understand your requirements fully) but the bullion dealers and metal
refiners who supply the jewellers. I don't know who they are in the US
or Australia, but here in the UK Johnson Matthey Metals would be the
first port of call. They can supply very pure Pt, Ir, Pd, Au, W, etc.

I don't fully understand your point about aluminium. If you are
wrapping it round tungsten wire, you are presumably then evaporating
it thermally rather than sputtering. So where does the argon come in?
Why don't you do this under ultra-high vacuum?

Best wishes
Chris

----- Original Message -----
} From: "Rosemary White" {rosemary.white-at-csiro.au}
To: "Tindall, Randy D." {TindallR-at-missouri.edu}
Cc: {microscopy-at-sparc5.microscopy.com}
Sent: Friday, June 14, 2002 12:48 AM


Hi Jerzy,

also look for potassium! Cells contain around 140 mM of K, which should translate to something in the range of 10 percent (biologists please correct me if I'm wrong) K atoms in the dry mass. Much more than Ca anyway.

Regards,
Heiko

------------------------------------------

Dr. Heiko Stegmann
AMD Saxony Manufacturing GmbH
Materials Analysis/TEM Group
Wilschdorfer Landstr. 101
01109 Dresden
Germany
Phone +49-351-2774167
Fax +49-351-2774199
e-mail heiko.stegmann-at-amd.com



Von: "jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com
[mailto:"jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com]
Gesendet: Freitag, 14. Juni 2002 06:23
An: Microscopy-at-sparc5.microscopy.com
Betreff: Bacteria on semiconductor


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Experts,

as a materials scientist I don't have too much exposure to bio-elemental analysis. Could someone in Bio-fields help out. What elements should I expect to find in biological residue? I am evaluating suspected bacterial deposits (contamination) and finding O, C, and Ca. Would there be any other elements present after the residue was encapsulated in high temperature deposited silicon oxide?

Thank you in advance.

Warmest Regards from Texas.

Jerzy



******************************************************
Jerzy Gazda, Ph.D. Advanced Micro Devices
Supervising Engineer 5204 E. Ben White Blvd. - MS 613
PCAL - AIM Section Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453 FAX: (512) 602-7470
jerzy.gazda-at-amd.com
******************************************************











From daemon Fri Jun 14 04:41:18 2002



From: Chris Jeffree :      cjeffree-at-srv0.bio.ed.ac.uk
Date: Fri, 14 Jun 2002 10:32:39 +0000
Subject: Encapsulite safelights

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All

I have been asked to specify safelights for a new darkroom and
TEM room. The old darkroom had a red Encapsulite fluorescent
strip, that was great for paper printing but we found that it was not
safe with EM Film. Is there an Encapsulite filter type you can
confidently recommend for use with Kodak SO163 EM Film, or
would you prefer something else? If so, what? The darkroom will
be 3x3metres, and ideally I want a diffuse, general room lighting
that will serve for both EM film loading and processing and bromide
paper printing.

Chris
==========================================
Dr. Chris Jeffree
University of Edinburgh
BIOSEM - Biological Sciences Electron Microscope Facility
Institute of Cell and Molecular Biology
Waddington Building
King's Buildings, Mayfield Road
EDINBURGH, EH9 3JN, Scotland, UK
Tel. #44 (0) 131 650 5554
FAX. #44 (0) 131 650 6563
Mobile 07710 585 401
email c.jeffree-at-ed.ac.uk
=========================================


From daemon Fri Jun 14 08:00:08 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Fri, 14 Jun 2002 08:55:13 -0400
Subject: Bacteria on semiconductor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jerzy;

I also deal in semiconductors, specifically failure analysis of same. On
occasion, I have to identify what appears to be an organic contaminant on
the surface or embedded into a structure such as a dielectric film, SiO2,
SiN etc. The tool that I find most useful for this is EDX [EDS] since these
contaminants generally detectable via SEM. Of course, the EDX system has
light element capability for O, C, etc.

More often than not, the elements you mentioned, C, O, Ca are present as
well as P, Cl, Mg, Fe. Also, many contaminants are simply skin flakes,
humans shed several million per day, which contain all of the above
elements. There is a reference to this in the Merck Index [composition of
skin/dermis/epidermis]. I can dig it out if you need it and it's a good
reference to substantiate or corroborate conclusions.

An additional nightmare in semiconductor contamination is cosmetics, e.g.
eye makeup, and that other powder stuff, I believe morticians also use it.
Since one cannot cover their eyes while looking through the eyepieces of a
microscope, little things like mascara break off of eyelashes and on to the
device. I think there is also a reference somewhere as to the elemental
composition of this stuff [makeup]. I'm sure most are proprietary
formulations like Helena Rubenstein's Stuff etc. but it's good to know
generally what's in these things.

The problem you may have with bacterial identification is the fact that
wafer processing uses high temperatures in deposition, annealing, plasma
treatment etc. At 750 C and higher, I doubt whether a bacteria will still
look like one if imaged in an SEM or optically. Imaging an organism is
probably the best identification of the "species", if that's the correct
term, if one can do so prior to all the high temp. processing.

I'm sure someone on the listserver, if not many, have great references on
SEM identification of bacteria.

Regards,

Peter Tomic
Team Leader
Analytical Dept.
Anadigics, Inc.





-----Original Message-----
} From: "jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com
[mailto:"jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com]
Sent: Friday, June 14, 2002 12:23 AM
To: Microscopy-at-sparc5.microscopy.com


Experts,

as a materials scientist I don't have too much exposure to bio-elemental
analysis. Could someone in Bio-fields help out. What elements should I
expect to find in biological residue? I am evaluating suspected bacterial
deposits (contamination) and finding O, C, and Ca. Would there be any other
elements present after the residue was encapsulated in high temperature
deposited silicon oxide?

Thank you in advance.

Warmest Regards from Texas.

Jerzy



******************************************************
Jerzy Gazda, Ph.D. Advanced Micro Devices
Supervising Engineer 5204 E. Ben White
Blvd. - MS 613
PCAL - AIM Section Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453 FAX: (512) 602-7470
jerzy.gazda-at-amd.com
******************************************************





From daemon Fri Jun 14 08:22:01 2002



From: Simon Dumbill :      simon.dumbill-at-aeat.co.uk (by way of
Date: Fri, 14 Jun 2002 08:12:00 -0500
Subject: TEM - SF6 disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Content-Type: text/html; charset=ISO-8859-1
Content-Description: HTML


We're in the process of planning the decommissioning of an EM430. Our
Safety and Environment people are not at all happy about the idea of
us just dumping the SF6 from the HT tank and, so far, we've not been
able to identify another disposal route. We've contacted refrigerant
disposal companies but SF6 is a new one to them and Air Products were
no use either. So if anyone out there has any experience of SF6
disposal I'd be very interested to hear from you.

Thanks,

Simon

{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}

Dr Simon Dumbill
Team Leader, Microstructural Characterisation
AEA Technology Nuclear Science
B14, Windscale
Seascale
Cumbria CA20 1PF

Tel: +44 (0)19467 72235
Fax: +44 (0)19467 72606

Email: {mailto:Simon.Dumbill-at-aeat.co.uk} Simon.Dumbill-at-aeat.co.uk

{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}


***********************************************************************
This transmission contains information which may be confidential and
which may also be privileged. It is intended for the named addressee
only. Unless you are the named addressee, or authorised to receive it
on behalf of the addressee you may not copy or use it, or disclose it
to anyone else. If you have received this transmission in error please
contact the sender. Thank you for your cooperation.
***********************************************************************

For more information about AEA Technology please visit our website at
http://www.aeat.co.uk

AEA Technology plc registered office 329 Harwell, Didcot, Oxfordshire OX11 0QJ.
Registered in England and Wales, number 3095862.


From daemon Fri Jun 14 09:58:49 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Fri, 14 Jun 2002 10:51:52 -0400
Subject: Re: Bacteria on semiconductor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Listers;

If anyone needs an EDX quantification of skin flakes, I have some on file.
However, there are so many variables in the quantification that I'm
uncertain if it would be useful. P & K are certainly among the elements of
skin.

Peter Tomic

-----Original Message-----
} From: "heiko.stegmann-at-amd.com"-at-sparc5.microscopy.com
[mailto:"heiko.stegmann-at-amd.com"-at-sparc5.microscopy.com]
Sent: Friday, June 14, 2002 3:48 AM
To: Microscopy-at-sparc5.microscopy.com


Hi Jerzy,

also look for potassium! Cells contain around 140 mM of K, which should
translate to something in the range of 10 percent (biologists please correct
me if I'm wrong) K atoms in the dry mass. Much more than Ca anyway.

Regards,
Heiko

------------------------------------------

Dr. Heiko Stegmann
AMD Saxony Manufacturing GmbH
Materials Analysis/TEM Group
Wilschdorfer Landstr. 101
01109 Dresden
Germany
Phone +49-351-2774167
Fax +49-351-2774199
e-mail heiko.stegmann-at-amd.com



Von: "jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com
[mailto:"jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com]
Gesendet: Freitag, 14. Juni 2002 06:23
An: Microscopy-at-sparc5.microscopy.com
Betreff: Bacteria on semiconductor


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Experts,

as a materials scientist I don't have too much exposure to bio-elemental
analysis. Could someone in Bio-fields help out. What elements should I
expect to find in biological residue? I am evaluating suspected bacterial
deposits (contamination) and finding O, C, and Ca. Would there be any other
elements present after the residue was encapsulated in high temperature
deposited silicon oxide?

Thank you in advance.

Warmest Regards from Texas.

Jerzy



******************************************************
Jerzy Gazda, Ph.D. Advanced Micro Devices
Supervising Engineer 5204 E. Ben White
Blvd. - MS 613
PCAL - AIM Section Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453 FAX: (512) 602-7470
jerzy.gazda-at-amd.com
******************************************************











From daemon Fri Jun 14 10:38:34 2002



From: Gib Ahlstrand :      giba-at-puccini.cdl.umn.edu
Date: Fri, 14 Jun 2002 10:34:02 -0500
Subject: Re: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Geetings Randy, and to others on this "thread":

} We have an Emitech cryopreparation unit associated with our Hitachi S4700
} FESEM.

snip!

} Also, if anybody has any wisdom to share on avoiding getting ice on samples
} during cryo runs, I'd love to hear it. We often (usually) get ice during
} transfers and need to sublime off the moisture before coating, otherwise the
} coating forms quasi-permanent ice "casts" that render the sample unusable.
} We've replaced/checked all the o-rings we can reach, so any special hints are
} very welcome.

On my Hitachi S3500N SEM, I have the Emitech K1150 cryo-prep unit, the
"little sibling" to the K1250, which I'll bet is the one you have? Mine
mounts onto the SEM sample chamber and consists of has air lock transfer
into a sputter coating chamber, then on into the SEM. The SEM vacuum system
is used for this unit, and I have to bleed the argon through the entire SEM,
as well as the sputter coater, but it works fine that way. The LN
cryo-freezing unit is a small seperate bench-top device, so I do need to
make a transfer from that, in room air, into the SEM mounted sputter coater
& air lock transfer unit. My system comes with a metal frostproofing shroud
which is picked up at the bottom of the LN freezing unit immediately after
the freeze and it tightly surrounds the sample and its stub in a nitrogen
gas atmosphere and prevents room air and humidity from getting to sample, so
I almost never see frost at the low to moderate mags we typically use. I
would think your system would also have such protection, so I wonder if
maybe your samples have surface moisture that looks like frost after
freezing?

In the event of frost contamination, here is a cheap trick that I use to
sublimate it off the surface once its in the SEM: Just hook up a bottle of
dry nitrogen gas to the air input of your SEM, via a regulator set to about
3-8 psi. I you already use N gas to vent your SEM, you're all set! Vent to
atmosphere and then immediately pump back down. All the while the cryostage
is kept at usual maximum cold temperature, about -120 C in my case, but the
gas imparts just enough thermal energy to the frost to sublimate it off. You
may not even have to vent all the way to room pressure before pumping back
down - experiment and see what works.

The big advantage is this only takes me 4 minutes for complete cycle.
Disadvantage is you must turn the beam off so you can't watch, but not a
problem. I don't like to defrost by heating up the stage and the entire mass
of the sample just to get some heat to the surface frost, as it takes so
long to do that plus cool down again, and you run the risk of overheating
and drying out the surface of the sample. Also with the K1150, because the
sputter coater is mounted to the SEM, if I should need to resputter after a
defrost, I just pull it back into the sputter chamber and there is no
transfer in air involved. Again, its rare that I have to do this now, on my
current system. Before, on my old Philips 500 SEM, I had to do air transfers
with NO frost shroud, so I did this N gas defrost more often.

Good luck!

Gib

} Thanks!
}
} Randy
}
} Randy Tindall
} EM Specialist
} Electron Microscopy Core Facility
} W122 Veterinary Medicine
} University of Missouri
} Columbia, MO 65211
} Tel: (573) 882-8304
} Fax: (573) 884-5414
} Email: tindallr-at-missouri.edu
} Web: http://www.biotech.missouri.edu/emc/
-- --------------------------
Gib Ahlstrand
Electron Optical Facility, University of Minnesota, CBS Imaging Center,
35 Snyder Hall, St. Paul, MN. USA. 55108 (612)624-3454
(612)625-5754 FAX, giba-at-puccini.cdl.umn.edu
http://www.cbs.umn.edu/ic/




From daemon Fri Jun 14 10:49:50 2002



From: Gib Ahlstrand :      giba-at-puccini.cdl.umn.edu
Date: Fri, 14 Jun 2002 10:46:08 -0500
Subject: Re: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I forgot to add this reference for my nitrogen gas defrost to my recent
posting on this topic. The pictures show an example of the results obtained:

Low-temperature low-voltage scanning electron microscopy of uncoated frozen
biological materials: A simple alternative. 1996. Microscopy &
Microanalysis. pp 918-19. (Minneapolis MSA meeting).

Reprints are still available!! (plug plug).

Gib

--
Gib Ahlstrand
Electron Optical Facility, University of Minnesota, CBS Imaging Center,
35 Snyder Hall, St. Paul, MN. USA. 55108 (612)624-3454
(612)625-5754 FAX, giba-at-puccini.cdl.umn.edu
http://www.cbs.umn.edu/ic/




From daemon Fri Jun 14 12:04:58 2002



From: Gordon Vrololjak :      gvrdolja-at-nature.Berkeley.EDU
Date: Fri, 14 Jun 2002 09:56:27 -0700 (PDT)
Subject: Re: TEM - SF6 disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think recycling SF6 through a company that does such work is the best
option.

Here is an article about SF6 I found useful:

Many people have heard of the environmental impact of greenhouse gas
emissions, such as chlorofluorocarbons. As a result, we have seen design
modifications in product packaging, such as changing aerosol propellants
to pumps and other environmentally safe dispensers. The IEEE has also been
concerned with these greenhouse gases, and in particular with sulfur
hexafluoride (SF6), which is commonly used as a dielectric and insulator
in circuit breakers, gas-insulated substations (GIS), and related
equipment for electrical transmission and distribution systems. According
to the Environmental Protection Agency (EPA), SF6 is a desirable gas
because of its stability when used in these electrical systems, but it has
undesirable properties when released into the environment. It has a long
life of about 3200 years, according to the estimates of the
Intergovernment-al Panel on Climate Change (IPCC). It is one of the worst
offenders of the greenhouse gases, having the highest global warming
potential.

Currently, there are no regulations regarding the disposal of SF6.
However, IEEE standards are encouraging manufacturers to safeguard the
release of SF6 into the air. IEEE Std C37.122.1-1993, a guide to GIS,
recommends against the unnecessary release of SF6 into the atmosphere
because of environmental concerns. Draft standard IEEE P1403, which
compares air-insulated substations and GIS, mentions that recent advances
in GIS construction include sophisticated equipment needed to reprocess
SF6. This recommendation means that less SF6 is released into the
atmosphere, helping to alleviate environmental concerns over the release
of greenhouse gases.

The EPA also has been involved with the subject of SF6. In August 1995, it
hosted a conference entitled Electrical Transmission and Distribution
Systems--Sulfur Hexafluoride and the Atmospheric Effects of Greenhouse Gas
Emissions. This international meeting included IEEE members as attendees.
The session urged voluntary compliance among users of SF6 so that further
official regulation is not needed in this area. This session also helped
to raise the visibility of SF6 disposal as a growing area of environmental
concern. Expect to see more discussion of SF6 disposal in the future.

For the final proceedings of the EPA greenhouse gas emissions conference,
contact the Atmospheric Pollution Prevention Division; US EPA 6202J;
Washington, DC 20460.


\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\
Gordon Ante Vrdoljak Electron Microscope Lab
ICQ 23243541 http://nature.berkeley.edu/~gvrdolja 26 Giannini Hall
gvrdolja-at-nature.berkeley.edu UC Berkeley
phone (510) 642-2085 Berkeley CA 94720-3330
fax (510) 643-6207 cell (510) 290-6793

On Fri, 14 Jun 2002, Simon Dumbill wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Content-Type: text/html; charset=ISO-8859-1
} Content-Description: HTML
}
}
} We're in the process of planning the decommissioning of an EM430. Our
} Safety and Environment people are not at all happy about the idea of
} us just dumping the SF6 from the HT tank and, so far, we've not been
} able to identify another disposal route. We've contacted refrigerant
} disposal companies but SF6 is a new one to them and Air Products were
} no use either. So if anyone out there has any experience of SF6
} disposal I'd be very interested to hear from you.
}
} Thanks,
}
} Simon
}
} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}
}
} Dr Simon Dumbill
} Team Leader, Microstructural Characterisation
} AEA Technology Nuclear Science
} B14, Windscale
} Seascale
} Cumbria CA20 1PF
}
} Tel: +44 (0)19467 72235
} Fax: +44 (0)19467 72606
}
} Email: {mailto:Simon.Dumbill-at-aeat.co.uk} Simon.Dumbill-at-aeat.co.uk
}
} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}
}
}
} ***********************************************************************
} This transmission contains information which may be confidential and
} which may also be privileged. It is intended for the named addressee
} only. Unless you are the named addressee, or authorised to receive it
} on behalf of the addressee you may not copy or use it, or disclose it
} to anyone else. If you have received this transmission in error please
} contact the sender. Thank you for your cooperation.
} ***********************************************************************
}
} For more information about AEA Technology please visit our website at
} http://www.aeat.co.uk
}
} AEA Technology plc registered office 329 Harwell, Didcot, Oxfordshire OX11 0QJ.
} Registered in England and Wales, number 3095862.
}



From daemon Fri Jun 14 12:38:38 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Fri, 14 Jun 2002 11:38:33 -0600
Subject: Bacteria on semiconductor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jerzy,

There are two other sources (probably more) for the contamination besides
bacteria or the materials that Peter mentioned:

1) Algae. When I worked on semiconductors, we had a clean room where some
parts of the DI Water lines used transparent hoses. We found algae growing
in the DI water.

2) The elements you mention (O, C, Ca) can also come from Calcium carbonate.
I had at one time some dirt on my wafers that included the same materials.
The dirt consisted of roundish contamination areas, some with a ring-like
structure. It turned out that we had slightly hard water and we did not dry
our materials carefully enough to avoid small droplets remaining on the
material. Further heat treatment then evaporated the water and left the
residue.

mike


} } } } } } } } } } WE HAVE MOVED { { { { { { { { {
please make a note of the new address below

Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: "jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com
[mailto:"jerzy.gazda-at-amd.com"-at-sparc5.microscopy.com]
Sent: Thursday, June 13, 2002 10:23 PM
To: Microscopy-at-sparc5.microscopy.com


Experts,

as a materials scientist I don't have too much exposure to bio-elemental
analysis. Could someone in Bio-fields help out. What elements should I
expect to find in biological residue? I am evaluating suspected bacterial
deposits (contamination) and finding O, C, and Ca. Would there be any other
elements present after the residue was encapsulated in high temperature
deposited silicon oxide?

Thank you in advance.

Warmest Regards from Texas.

Jerzy



******************************************************
Jerzy Gazda, Ph.D. Advanced Micro Devices
Supervising Engineer 5204 E. Ben White
Blvd. - MS 613
PCAL - AIM Section Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453 FAX: (512) 602-7470
jerzy.gazda-at-amd.com
******************************************************





From daemon Fri Jun 14 13:25:53 2002



From: sghoshro-at-NMSU.Edu
Date: Fri, 14 Jun 2002 12:19:36 -0600 (MDT)
Subject: Re: Encapsulite safelights

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chris,

We use Kodak light amber safelight for both EM film and paper and we use
Kodak polycontrast printing paper and 4489 EM films. I am quite sure the
light amber is safe for SO-163 and bromide paper.

Best wishes,

Soumitra

*************************************************************
Soumitra Ghoshroy Ph.D.
Electron Microscopy Lab
Department of Biology
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
URL:http://confocal.nmsu.edu/eml

On Fri, 14 Jun 2002, Chris Jeffree wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear All
}
} I have been asked to specify safelights for a new darkroom and
} TEM room. The old darkroom had a red Encapsulite fluorescent
} strip, that was great for paper printing but we found that it was not
} safe with EM Film. Is there an Encapsulite filter type you can
} confidently recommend for use with Kodak SO163 EM Film, or
} would you prefer something else? If so, what? The darkroom will
} be 3x3metres, and ideally I want a diffuse, general room lighting
} that will serve for both EM film loading and processing and bromide
} paper printing.
}
} Chris
} ==========================================
} Dr. Chris Jeffree
} University of Edinburgh
} BIOSEM - Biological Sciences Electron Microscope Facility
} Institute of Cell and Molecular Biology
} Waddington Building
} King's Buildings, Mayfield Road
} EDINBURGH, EH9 3JN, Scotland, UK
} Tel. #44 (0) 131 650 5554
} FAX. #44 (0) 131 650 6563
} Mobile 07710 585 401
} email c.jeffree-at-ed.ac.uk
} =========================================
}
}



From daemon Fri Jun 14 14:05:54 2002



From: jerzy.gazda-at-amd.com
Date: Fri, 14 Jun 2002 13:59:16 -0500
Subject: Bacteria on semiconductor - thank you

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to all that have replayed to my question from last night.

For trade reasons I cannot discuss all the details of the defect I am investigating, but it is sufficient to say, we traced its formation to one of water based wet cleans. I was provided samples that had defects encapsulated into layers of Si-oxide and went through high temperature processes. Therefore the microorganism was burned up and resulting residue does not have structure (amorphous blob of darker contrast in oxide). Based on your replays, I will follow up with analysis on blanket Si wafers processed in the same sink and attempt to image the micro-organism with SEM. Hopefully the structure will be preserved. Elements listed in your replays are: H, C, O, K, Ca, Na, and if bacteria is metal reducing S, Fe, etc.

Again, thank you all for your help.

Even warmer Regards from Texas (104F).

Jerzy

******************************************************
Jerzy Gazda, Ph.D. Advanced Micro Devices
Supervising Engineer 5204 E. Ben White Blvd. - MS 613
PCAL - AIM Section Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453 FAX: (512) 602-7470
jerzy.gazda-at-amd.com
******************************************************





From daemon Fri Jun 14 15:38:03 2002



From: Gerroir, Paul J :      Paul.Gerroir-at-crt.xerox.com
Date: Fri, 14 Jun 2002 16:10:22 -0400
Subject: SEM - Use of Cyro-transfer/Sputter system. Fee for use

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good Afternoon,
I would like to hear from anyone who can help me locate a lab having a SEM
equipped with a cryo-transfer/sputter system. I am interested in examining
small (5 to 10 micron) polymer particles that can be fractured at low
temperature then coated and subsequently examined, preferably at low ( {5kV)
accelerating voltage. Hopefully such a lab/facility exists in Southern
Ontario, somewhere between Toronto and Hamilton would be great. Thanks.

Paul

Paul J. Gerroir
Microscopy
Materials Characterization
Xerox Research Centre of Canada
2660 Speakman Drive
Mississauga, Ontario L5K 2L1

Phone: (905) 823-7091, ext. 216
FAX: (905) 822-7022
email: paul.gerroir-at-crt.xerox.com



From daemon Fri Jun 14 20:00:27 2002



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Fri, 14 Jun 2002 17:51:58 -0700
Subject: RE: TEM - SF6 disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Simon;
Our service engineer piped it into a plastic garbage bag, which he
carried outside and dumped. This is relatively easy because SF6 is
considerably denser than air, and safer than dumping it in the room, since
the HVAC would have difficulty eliminating it. Our new labs have the SF6
bleed plumbed to scrubbed exhaust.

John Mardinly
Intel




Content-Type: text/html; charset=ISO-8859-1
Content-Description: HTML


We're in the process of planning the decommissioning of an EM430. Our
Safety and Environment people are not at all happy about the idea of
us just dumping the SF6 from the HT tank and, so far, we've not been
able to identify another disposal route. We've contacted refrigerant
disposal companies but SF6 is a new one to them and Air Products were
no use either. So if anyone out there has any experience of SF6
disposal I'd be very interested to hear from you.

Thanks,

Simon

{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}

Dr Simon Dumbill
Team Leader, Microstructural Characterisation
AEA Technology Nuclear Science
B14, Windscale
Seascale
Cumbria CA20 1PF

Tel: +44 (0)19467 72235
Fax: +44 (0)19467 72606

Email: {mailto:Simon.Dumbill-at-aeat.co.uk} Simon.Dumbill-at-aeat.co.uk

{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}


***********************************************************************
This transmission contains information which may be confidential and
which may also be privileged. It is intended for the named addressee
only. Unless you are the named addressee, or authorised to receive it
on behalf of the addressee you may not copy or use it, or disclose it
to anyone else. If you have received this transmission in error please
contact the sender. Thank you for your cooperation.
***********************************************************************

For more information about AEA Technology please visit our website at
http://www.aeat.co.uk

AEA Technology plc registered office 329 Harwell, Didcot, Oxfordshire OX11
0QJ.
Registered in England and Wales, number 3095862.


From daemon Sat Jun 15 01:53:52 2002



From: Bill & Sue Tivol :      wtivol-at-earthlink.net
Date: Fri, 14 Jun 2002 23:39:31 -0400
Subject: Re: TEM - SF6 disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


on 6/14/02 9:12 AM, Simon Dumbill (by way of MicroscopyListserver) at
simon.dumbill-at-aeat.co.uk wrote:

} We're in the process of planning the decommissioning of an EM430. Our
} Safety and Environment people are not at all happy about the idea of
} us just dumping the SF6 from the HT tank and, so far, we've not been
} able to identify another disposal route. We've contacted refrigerant
} disposal companies but SF6 is a new one to them and Air Products were
} no use either. So if anyone out there has any experience of SF6
} disposal I'd be very interested to hear from you.
}
} Thanks,
}
Dear Simon,
One thing to try is to freeze the SF6 into an empty gas bottle by
connecting the tank, regulator, etc. to a fitting on the HT tank, then
putting the bottom of the gas bottle in a bucket of dry ice (assuming that
the S&E people are not alarmed at the resulting CO2). At room temp, the SF6
will be mostly in the liquid state with some gas at 330 psi--its vapor
pressure. The SF6 is pretty valuable, so you might easily find someone to
take it off your hands. Good luck.
Yours,
Bill Tivol



From daemon Mon Jun 17 03:38:51 2002



From: Rosemary White :      rosemary.white-at-csiro.au
Date: Mon, 17 Jun 2002 18:21:02 +1000
Subject: Re: SEM/ Cryo unit sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chris,

Yes, we only use this freezing method for fairly rigid smallish plants and
insects or their parts. We call it slow-freezing because it's slower than
plunging in liquid ethane or propane (lab) or nitrogen (field) or snap
freezing between pliers cooled in LN2. But it takes less than a minute,
usually only a few seconds, for the tissue to freeze in the prep chamber,
so they don't lose too much surface water, not enough to damage the surface
structure, at least. Once it's frozen, it wouldn't lose more than a
plunge-frozen specimen, I shouldn't think. We don't deal with cultured
animal cells at all, for SEM at least, and I'd be wary of recommending this
method for soft and/or wet tissues (might be worth trying once, though).
And when we need to do analysis, the plunge frozen tissue is first planed
to give a fresh "clean" flat surface. Then we do need to get rid of frost
and also etch very lightly.

Re metals, Alfa Aeser is now controlled by Johnson Matthey, I think, and is
certainly our first port of call for other than gold, which we have been
able to get at highest purity from the local jeweller.

Sorry, brain disengaged re. Al coating - of course, this is evaporative
coating..... so we have to clean the chamber windows every now and then,
before they turn into mirrors.
cheers,
Rosemary



Rosemary
While I concede the slow-freeze method works on some specimens - it
can as you indicate be
particularly successful with leaves or small insects that are covered
with a cuticle -
such recidivism should only be practised by consenting adults fully
aware of the implications, which are
a) wet stuff (possibly including organic solvents) has to be
introduced to the high vacuum of the prep chamber thus contaminating
cold stage and anticontaminator of the prep chamber
b) uncontrolled evaporation from the surface of the un-fixed specimen
c) thus specimens without the benefit of a protective covering (such
as cultured cells) will show variable amounts of drying artefacts.
I suppose the frank reality is that all etching / drying procedures
result in artefacts, but there is a comforting illusion of control if
you can get the same artefacts with reasonable reproducibility!

On the topic of gold for targets, I was not recommending jewellers as
a source of metals (although that is fine if they, like yours, seem to
understand your requirements fully) but the bullion dealers and metal
refiners who supply the jewellers. I don't know who they are in the US
or Australia, but here in the UK Johnson Matthey Metals would be the
first port of call. They can supply very pure Pt, Ir, Pd, Au, W, etc.

I don't fully understand your point about aluminium. If you are
wrapping it round tungsten wire, you are presumably then evaporating
it thermally rather than sputtering. So where does the argon come in?
Why don't you do this under ultra-high vacuum?

Best wishes
Chris





From daemon Mon Jun 17 06:45:53 2002



From: George Laing :      scisales-at-ngscorp.com
Date: Mon, 17 Jun 2002 07:37:51 -0400
Subject: RE: Encapsulite safelights

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


According to Kodak, the correct safelight for SO-163 film is either
the GBX-2 or a #1(red). For 4489, the correct safelight filters are
either the OC(amber) or OA(greenish yellow).
For more safelight information see
http://www.kodak.com/US/en/health/scientific/products/electronmicrography/da
rkroom.shtml

Encapsulite has filters for almost every use. They also have a
dual lamp fixture for those needing two filters.
See http://www.encapsulite.com/doublelamp.htm If you contact them, they
can give you their equivalent to the GBX-2.


George

George Laing
National Graphic Supply
ph 800.223.7130 x3109
f 800.832.2205
email scisales-at-ngscorp.com
}
} Is there an Encapsulite filter type you can
} confidently recommend for use with Kodak SO163 EM Film, or
} would you prefer something else? If so, what? The darkroom will
} be 3x3metres, and ideally I want a diffuse, general room lighting
} that will serve for both EM film loading and processing and bromide
} paper printing.
}




From daemon Mon Jun 17 08:12:52 2002



From: Lesley S. Bechtold :      lsb-at-jax.org
Date: Mon, 17 Jun 2002 09:04:59 -0400
Subject: Looking for Jan Schwarz

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have her email address? She's located in Vermont.

Thanks!

Lesley Bechtold



Lesley S. Bechtold
Supervisor, Biological Imaging
The Jackson Laboratory
600 Main St.
Bar Harbor, ME 04609
207-288-6191



From daemon Mon Jun 17 12:27:26 2002



From: Mae :      pqqljxitg-at-cci.be
Date: Tue, 04 Jun 2002 08:47:29 +0900
Subject: Learn how to Outperform the Dow Jones industrial average

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Doubling your investment is more achievable
than you think!

200% return could have been achieved in less than 30 days!
Learn about our strategy trading in the InterBank Foreign
Currency Exchange.

Get our Reports, Charts and Strategies on the U.S. Dollar vs
Japanese Yen and Euro.

Example:

A $5,000 Investment in the Euro vs the dollar, "properly positioned",
on 7/17/01 could possibly have returned $12,500 on 08/10/01.

Must be 21 years of age.
For additional information, stop by our website..today.

http://builtit4unow.com/yen
===================================================================================
This email advertisement is sent out to those who subscribed
on our web site or one of our many affiliated web sites.
If you received this email in error or you would like to opt-out
from our database, please go to the following:

http://builtit4unow.com/options




From daemon Mon Jun 17 13:48:44 2002



From: hard-at-acsu.buffalo.edu
Date: Mon, 17 Jun 2002 14:38:17 -0500
Subject: Course Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Course Announcement

Title: Optical Microscopy and Imaging in the Biomedical Sciences

When: October 9 - October 18, 2002

Where: Marine Biology Laboratory, Woods Hole, MA, USA

Tuition: $2200 (Includes room and board, text, handouts, supplies)

Application Deadline: July 25, 2002

Admission application and information:
Carol Hamel, Admissions Coordinator
Marine Biological Laboratory
7 MBL Street
Woods Hole, MA 02543-1015
(508) 289-7401
Internet: admissions-at-mbl.edu
WWW: http://www.mbl.edu (Application forms available via Adobe Acrobat)

Course Director: Colin S. Izzard, State University of New York -at- Albany
Phone: [518] 442 - 4367
EMail: csizzard-at-csc.albany.edu

Course Description:

For Whom:
Designed primarily for research scientists, physicians, postdoctoral
trainees and advanced graduate students in animal, plant, medical and
material sciences. Non-biologists seeking a comprehensive introduction to
microscopy and video-imaging will benefit greatly from this course as well.
There are no specific prerequisites, but an understanding of the basic
principles of optics is desirable. Limited to 24 students.

The eight day course consists of lectures, laboratory demonstrations,
exercises and discussions that will enable the participant to obtain and
interpret microscope images of high quality, to perform quantitative optical

measurements, and to produce photographic and video records for
documentation
and analysis.

Topics to be covered include:
principles of microscope design and image formation
bright field, dark field, phase contrast, polarized light,
differential
interference contrast, interference reflection, and fluorescence
microscopy
confocal scanning microscopy, multiphoton excitation fluorescence
microscopy and image deconvolution
digital image restoration and 3-D reconstruction
video imaging, recording, enhancement, and intensification
analog and digital image processing and analysis
fluorescent probes and ratiometric-imaging
laser tweezers and laser scissors

Applications to live cells will be emphasized; other specimens will be
covered as well.

Students will have direct hands-on experience with state-of-the-art
microscopes, video cameras, recorders and image processing equipment
provided
by major optical and electronics companies. Instruction will be provided by
experienced staff from universities and industry.

STUDENTS ARE ENCOURAGED TO BRING THEIR OWN BIOLOGICAL (PRIMARY
CULTURES, CELL LINES, PREPARED SLIDES, ETC.) AND MATERIAL SPECIMENS AND TO
USE THEM FOR COURSE EXERCISES, WHERE APPROPRIATE. Cell culture facilities
are available. Students are highly encouraged to discuss individual research
problems with the academic and commercial faculty.

For faculty list and additional information, see: http://www.mbl.edu




From daemon Mon Jun 17 15:39:13 2002



From: mohammed y abdulrawoof / f40z006 75760 :      mdyousuf-at-KSU.EDU.SA
Date: Mon, 17 Jun 2002 23:23:45 -0300 (GMT)
Subject: Thanks and appreciation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,
This list server is of great help, when cornered into bottle- necks.
Thanks to every one who replied to my Osmium Tetroxide question. Thanks to
all the expert advise, which helped me in using the many ampules which
were just about to be discarded from the stores. My deep regards to
Nestor who keeps the show running.

Mohammed Yousuf Abdul-Rawoof
Dept. of Zoology
College of Science
King Saud University
POB 2455, Riyadh 11451
Saudi Arabia.

mdyousuf-at-ksu.edu.sa
mdyousuf99-at-hotmail.com



From daemon Tue Jun 18 08:29:26 2002



From: PHYSIOL4-at-AKAD.SUN.AC.ZA
Date: Tue, 18 Jun 2002 15:11:09 +0200
Subject: Thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all

Thanks for all your ideas and input, I am currently at the end stages
of attempting to fix sugarcane in a 4% paraformaldehyde, 0.5%
glutaraldehyde mixture (I think Jan Coetzee suggested I try that),
and will let you know how that went. Thanks also to Rosemary White
for that reference about in situs on non-embedded tissues, will
definitely check that out.

Cheers!
Gabrielle


Gabrielle Turner
Institute for Plant Biotechnology
University of Stellenbosch
7600

Cell: 083 324 7453


From daemon Tue Jun 18 16:56:35 2002



From: Chun-Ming Li :      lichun-at-uiuc.edu
Date: Tue, 18 Jun 2002 16:41:41 -0500
Subject: TEM sample preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear friends:
I have a problem in preparing a TEM sample (plan view). The
as-received sample is as follow:
On a soda-lime glass (thickness 2-4 mm) substrate, deposit a
molybdenum layer (thickness 0.1-1.5 micron), and then deposit a layer of
CuInSe2 (thickness 0.7-2.5 micron).
After dissolved the glass with HF (hydrofluoric) acid (cover the layers
with lacquer before etching, otherwise the multilayers will also be
dissolved), I immersed the multilayers into acetone. But the layers are
too brittle that they broke into tiny chips. Even sometimes I can get a
big chip, it is still very difficult to fix it on a grid, since it is
too brittle.
I would appreciate it very much if you can give me any suggestions
to solve this problem, i.e., how to get a plan-view TEM sample from the
as-received sample. The major problem would be how to get rid of the
glass and fix the multilayers to grid before ion milling.
Thank you in advance!

Chun-Ming Li, Ph.D.
Dept. of Mater. Sci. and Engi.
University of Illinois at Urbana-Champaign




From daemon Tue Jun 18 23:00:26 2002



From: Walck, Scott D. :      walck-at-ppg.com
Date: Tue, 18 Jun 2002 23:35:50 -0400
Subject: TEM sample preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You do not have to dissolve the glass. You can get the layers off by suspending the sample in the HF vapors for a while with plastic tongs and then gently floating the samples off in water. A small amount of HF in a plastic beaker will work fine. You should have small chips to work with so that the HF vapors do not have to penetrate far. You can then float the films off in water much like you do with films on salt crystals. For better support, use a high mesh grid, 400 of more. You can also get a little better support with holey carbon support films on the grids. Once the samples are on the grids, you will have to dissolve the lacquer. For picking the films up from the water, you should try using a wire loop as it will be more gentle on your fragile samples than trying to pick them up on the grids directly. Electron Microscopy Science sells the "Perfect Loop" which might work well for your application.

One issue you have though is that at 1.5 um, the samples will not be electron transparent, but I assume that you have thought of that.

Be very careful with HF! Do not let it come in contact with your skin and do not breath the vapors!



-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8515 (fax)



-----Original Message-----
} From: Chun-Ming Li [mailto:lichun-at-uiuc.edu]
Sent: Tuesday, June 18, 2002 5:42 PM
To: Microscopy-at-sparc5.microscopy.com


Dear friends:
I have a problem in preparing a TEM sample (plan view). The
as-received sample is as follow:
On a soda-lime glass (thickness 2-4 mm) substrate, deposit a
molybdenum layer (thickness 0.1-1.5 micron), and then deposit a layer of
CuInSe2 (thickness 0.7-2.5 micron).
After dissolved the glass with HF (hydrofluoric) acid (cover the layers
with lacquer before etching, otherwise the multilayers will also be
dissolved), I immersed the multilayers into acetone. But the layers are
too brittle that they broke into tiny chips. Even sometimes I can get a
big chip, it is still very difficult to fix it on a grid, since it is
too brittle.
I would appreciate it very much if you can give me any suggestions
to solve this problem, i.e., how to get a plan-view TEM sample from the
as-received sample. The major problem would be how to get rid of the
glass and fix the multilayers to grid before ion milling.
Thank you in advance!

Chun-Ming Li, Ph.D.
Dept. of Mater. Sci. and Engi.
University of Illinois at Urbana-Champaign




From daemon Wed Jun 19 02:47:38 2002



From: Richard Beanland :      richard.beanland-at-bookham.com
Date: Wed, 19 Jun 2002 08:38:18 +0100
Subject: TEM sample preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chun-Ming,
One way to pick up small thin fragments is to use a mesh grid with a
thin coating of epoxy. I use Devcon 5-minute epoxy. Choose a mesh grid
with holes smaller than your typical fragment size. Place it on filter
paper and cover it with a blob of glue. Then take another piece of filter
paper and press down on top to remove some glue. Separate the papers, move
the grid to a clean region of paper, and repeat 2 or 3 times. This will
give you a 'tacky' mesh grid with no epoxy covering the holes. You can then
use this to pick up fragments of film.
When looking at thin films on glass I have usually tried to make the
amount of glass to be dissolved as small as possible by grinding from the
back to a thickness of a grade zero cover slip (~120 um). You can then
dissolve the glass in a small amount of HF, which can be diluted massively
when the glass is gone, and filtered through coarse filter paper to collect
the fragments. When the paper is dry you can use a tacky grid to pick up
the fragments. In your case, I guess you would take the lacquer sheet (with
film attached) into a solvent and filter the solvent when the lacquer is
dissolved. Again massive dilution may help, in this case to reduce the
amount of lacquer which dries on the fragments.

Good luck,


Richard Beanland

__________________________
Richard Beanland
Analytical Services,
Bookham Technology plc.,
Caswell,
Towcester,
Northants NN12 8EQ
UK
e-mail richard.beanland-at-bookham.com



-----Original Message-----
} From: Chun-Ming Li [mailto:lichun-at-uiuc.edu]
Sent: 18 June 2002 22:42
To: Microscopy-at-sparc5.microscopy.com


Dear friends:
I have a problem in preparing a TEM sample (plan view). The
as-received sample is as follow:
On a soda-lime glass (thickness 2-4 mm) substrate, deposit a
molybdenum layer (thickness 0.1-1.5 micron), and then deposit a layer of
CuInSe2 (thickness 0.7-2.5 micron).
After dissolved the glass with HF (hydrofluoric) acid (cover the layers
with lacquer before etching, otherwise the multilayers will also be
dissolved), I immersed the multilayers into acetone. But the layers are
too brittle that they broke into tiny chips. Even sometimes I can get a
big chip, it is still very difficult to fix it on a grid, since it is
too brittle.
I would appreciate it very much if you can give me any suggestions
to solve this problem, i.e., how to get a plan-view TEM sample from the
as-received sample. The major problem would be how to get rid of the
glass and fix the multilayers to grid before ion milling.
Thank you in advance!

Chun-Ming Li, Ph.D.
Dept. of Mater. Sci. and Engi.
University of Illinois at Urbana-Champaign




=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.

No part of this message can be considered a request for goods or
services.
=======================================================================
Any questions about Bookham's E-Mail service should be directed to postmaster-at-bookham.com.



From daemon Wed Jun 19 06:18:31 2002



From: Sousan Abolhassani :      sousan.abolhassani-at-psi.ch
Date: Wed, 19 Jun 2002 13:07:37 +0200
Subject: Re: TEM sample preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chun-Ming,

As your multi layer is thick, in any case you should thin it
afterwards.
Instead of going through the problem of dissolving and collecting
such a layer that you still need to thin by other techniques
there are at least three possibilities (that do not need suffisticated
facilities):

a. If you want to see the top view of your multi-layer, then
it means that you should thin your sample from the molybdenum
side until it is electron transparent. For this, the sample
should be grinded from the glass side to a thickness of the order
of 20-50 microns, then mounted on a ring sample holder used
for ion milling. (If you can grind down to 30 micrometers it needs a
shorter time for ion-milling, but if you cann't it is not important).
Then you ion-mill your sample from the glass side, until you see a hole
and you know that the thinnest part is in the CuInSe2 region.

Note that in this case you will not see the molybdenum, because your
CuInSe2 is 0.7-2.5 micrometers as you say.
If you wish to observe the molybdenum as well, you could take another
sample and proceed as above, but stop milling when the sample is
near the Mo (you could see from colour changes) and then ion mill
from the other side.

b. The second possibility is to do a tripod polishing and this can
be done with different orientations. All layers will be observable
and from my experience this is the most convenient method, that
does not need to complicated facilities.

c. The last method, that I use for quick determinations of the
material before the above elaborate methods, is to scratch the
surface to be observed by a diamond pointer and then put the
sctratched debris in a solvent that does not attack the film
(you should use a very small recipient not to loose your material)
and collect the debris on a grid with a thin support film.
This method is very rapid and can quickly bring some information
before you proceed with more elaborate techniques.


Hope this helps,

Sousan



E-mail: sousan.abolhassani-at-psi.ch
http://www.psi.ch


Chun-Ming Li wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Dear friends:
} I have a problem in preparing a TEM sample (plan view). The
} as-received sample is as follow:
} On a soda-lime glass (thickness 2-4 mm) substrate, deposit a
} molybdenum layer (thickness 0.1-1.5 micron), and then deposit a layer of
} CuInSe2 (thickness 0.7-2.5 micron).
} After dissolved the glass with HF (hydrofluoric) acid (cover the layers
} with lacquer before etching, otherwise the multilayers will also be
} dissolved), I immersed the multilayers into acetone. But the layers are
} too brittle that they broke into tiny chips. Even sometimes I can get a
} big chip, it is still very difficult to fix it on a grid, since it is
} too brittle.
} I would appreciate it very much if you can give me any suggestions
} to solve this problem, i.e., how to get a plan-view TEM sample from the
} as-received sample. The major problem would be how to get rid of the
} glass and fix the multilayers to grid before ion milling.
} Thank you in advance!
}
} Chun-Ming Li, Ph.D.
} Dept. of Mater. Sci. and Engi.
} University of Illinois at Urbana-Champaign


From daemon Wed Jun 19 07:26:37 2002



From: L. D. Marks :      ldm-at-risc4.numis.nwu.edu
Date: Wed, 19 Jun 2002 07:23:38 -0500 (CDT)
Subject: PEELS on a PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a PC compatible interface for a Gatan
PEELS that they would be willing to donate (or sell) to
a worthy cause? I would also be interested in a Mac version,
although I would prefer a PC.

-----------------------------------------------
Laurence Marks
Department of Materials Science and Engineering
MSE Rm 2036 Cook Hall
2225 N Campus Drive
Northwestern University
Evanston, IL 60201, USA
Tel: (847) 491-3996 Fax: (847) 491-7820
mailto:ldm-at-risc4.numis.nwu.edu
http://www.numis.nwu.edu
-----------------------------------------------




From daemon Wed Jun 19 07:57:45 2002



From: Gabriel A. Rosa :      micros-at-bg.fcen.uba.ar
Date: Wed, 19 Jun 2002 09:51:01 -0300
Subject: I need help (spectrofluorometers)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi listers,
My name is Gabriel, and in my University (of Buenos Aires) are two
spectrofluorometers (one aminco SPF500 and one Perkin Elmer 650-40) both
without electric source for the lamps.
Anything have an electric source for this equipment for donation?
Or anything have the electrical chart for construct this??

T.I.A.

Gabriel Adriano Rosa
Centro de Imagenes y Microscopia
Departamento de Ciencias Biologicas
Facultad de Ciencias Exactas y Naturales, Universidad de Buenos Aires
Ciudad Universitaria, 4 piso, Pab. II,
CP 1428, Buenos Aires, ARGENTINA
TE -(54-11)-4576-3349
FAX (54-11)-4576-3384
e-mail cimic-at-bg.fcen.uba.ar






From daemon Wed Jun 19 09:49:41 2002



From: John Mansfield :      jfmjfm-at-umich.edu
Date: Wed, 19 Jun 2002 10:35:20 -0400
Subject: Gatan DuoMill parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi there,

We are still running a 14+ year old Gatan DuoMill and are having a few
problems with one of the Whisperloks leaking a little. There is a
Swagelock seal on the bottom of the whole assembly where the braid attaches
for the liquid nitrogen dewar for cold milling. This seal is so tightly
bound to the aluminum rod that runs through the center of the whole thing
that we cannot take the Swagelock stuff off to try and replace the ferrule
or the aluminum rod to try and improve the seal.

My question is, does anyone have a DuoMill that they have retired that we
could buy/scrounge parts from to try and fix this?

I can go to Gatan and buy a new part but I think that they may have to have
them made since they no longer make the DuoMill and it seems they don't have
this part in stock. Having one made will probably make this rather
expensive and I am trying to do this on the cheap. Can anyone be of
assistance?

ADVthanksANCE.

Jfm

--
John Mansfield PhD MInstP
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (734) 936-3352 FAX (734) 763-2282 Cell. Phone: (734) 834-3913
(Leaving a phone message at 936-3352 is preferable to 834-3913)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html
Location: Lat. 42! 16' 48" Long. 83! 43' 48"




From daemon Wed Jun 19 13:28:46 2002



From: Yan Xin :      xin-at-magnet.fsu.edu
Date: Wed, 19 Jun 2002 14:21:51 -0400
Subject: Jet polishing recipe for stainless steel 316LN

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I need to make TEM samples from the bulk stainless steel.
I would appreciate very much if anyone could give me the recipe to make TEM
samples of stainless steel 316LN using jet polishing.

Thanks

Regards
Yan xin
=======================================
Yan Xin (Ph.D)
Magnet Science & Technology
National High Magnetic Field Laboratory
Florida State University
1800 E. Paul Dirac Drive
Tallahassee, FL 32310
Tel: (850) 644 1529
Fax: (850) 644 0867
========================================





From daemon Wed Jun 19 15:33:25 2002



From: Krzysztof Herman :      kherman-at-labsoft.com.pl
Date: Wed, 19 Jun 2002 22:26:37 +0200
Subject: Philips EM420 IGP Power Supply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello List

Does somebody have the "IGP power supply" (or maybe some of it's components) to sell cheap
or even donate to Czech microscopists ??

(I don't count on many offers ;-)) ... but to ask is allways worthwise..

regards

Krzysztof Herman
Labsoft, ul.Bażancia 45A
02-892 Warszawa, Poland
tel/fx: (+48 22)6449753, 6449750
mobile: (+48 601)307456





From daemon Wed Jun 19 16:33:25 2002



From: Dmrelion-at-aol.com
Date: Wed, 19 Jun 2002 17:26:19 EDT
Subject: immersion oil removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear list members,

One of my associates has a large number of thin sections of limestone that he
wants to look at with cathodoluminescence (CL). The problem is that the thin
sections have been coated already with immersion oil for other microscopy
observations and this may create problems with the vacuum and electron beam
in the CL apparatus. Does anyone have suggestions on the best/easiest way to
remove the immersion oil? The oil is identified as "Immersion oil for
Microscopy, Cargille Type A, Formula Code 1248" and is said to be made up of
synthetic hydrocarbons and natural petroleum derivatives.

He will appreciate any and all suggestions.

Thank you.

Don Marshall

RELION Industries
PO Box 12
Bedford, MA 01730

781-275-4695 (Phone)
781-271-0252 (FAX)

dmrelion-at-aol.com

cathodoluminescence, mass spectroscopy, electron beam technology

http://www.excitingelectrons.com


From daemon Wed Jun 19 16:38:23 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Wed, 19 Jun 2002 17:31:45 -0400
Subject: FW: Darkroom safelights

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In spite of the fact that I think darkrooms are going to disappear either
because Kodak and friends go the way of Polaroid or because the digital
detectors get so good that they equal (though I can't see how!!!) the
resolution of film, I am following Chris' suggestion to put my response to
his safelight question on the list.

I have done this search before, so it wasn't just for his benefit. I
especially liked my find of the Butzi review which included some info about
LCD safelights which I fortuitously found again in the next to last
reference I gave.

Well, here it is, and I hope it's useful for others. I already have my
Thomas and I have the spare bulb. Our 20' x 15' of darkroom is completely
ready for all the students who will be using our new FEI Tachni 12T with its
Gatan 1000 side-mounted CCD camera. My analysis of the cost comparison for
a class of twenty between the CCD and Kodak 4489 has the Space Committee
already salivating about unused space. Unfortunately, the Polymer Chemists
are insisting on 4489 as are the Physicists. So, I'm ordering a 5' round
table so the rest of us can play poker while the few hard-timers are taking
pictures on the scope. they will be able to process during our breaks. I
am also generating a complete description of the original method of
agitation that was replaced by nitrogen burst technology which is no longer
cost effective at all.

Regards to all,

Fred



} ----------
} From: Monson, Frederick C.
} Sent: Wednesday, June 19, 2002 12:32 PM
} To: 'c.jeffree-at-ed.ac.uk'
} Subject: RE: Encapsulite safelights
}
} Morning Chris,
} I have been using a Thomas Super Duplex (~$200) for quite a long
} time, and I have noticed no problems with it an Kodak 4489. I have not
} tried it with SO163 or brands other than Kodak. I give you a site on
} which it is critiqued: http://www.butzi.net/reviews/darkroom.htm.
}
} Kodak Darkroom hints are very helpful at design time.
} http://www.kodak.com/cluster/global/en/consumer/products/techInfo/k4/k4Pla
} cement.shtml
}
} Safelight Recommendations - note the Table link
} http://www.kodak.com/global/en/consumer/products/techInfo/k4/k4Recomnd.sht
} ml#93831
}
} And more:
} http://www.silverprint.co.uk/dark19.html
}
} Ted Pella's ~$200 Duplex for $300
} http://www.tedpella.com/photo_html/photo10.htm
}
} LED Safelights:
} http://www.barbierielectronic.com/products/Safelights_en.htm
}
} Thomas at B&H
} http://198.77.206.55/links/534.html Choose {Thomas DUB ....} to get to
} good info on Thomas systems. Really hard to get good info on this stuff.
}
} Hope this helps,
}
} Regards,
}
} Fred
}
} Frederick C. Monson, PhD
} Center for Advanced Scientific Imaging
} Schmucker II Science Center
} West Chester University
} South Church Street and Rosedale
} West Chester, Pennsylvania, USA, 19383
} Phone: 610-738-0437
} FAX: 610-738-0437
} fmonson-at-wcupa.edu
} CASI URL: http://darwin.wcupa.edu/casi/
} WCUPA URL: http://www.wcupa.edu/
} Visitors URL: http://www.wcupa.edu/_visitors/
}
}
}
}
}
}
}
}
} ----------
} From: Chris Jeffree
} Reply To: c.jeffree-at-ed.ac.uk
} Sent: Friday, June 14, 2002 6:32 AM
} To: microscopy-at-sparc5.microscopy.com
} Subject: Encapsulite safelights
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear All
}
} I have been asked to specify safelights for a new darkroom and
} TEM room. The old darkroom had a red Encapsulite fluorescent
} strip, that was great for paper printing but we found that it was not
} safe with EM Film. Is there an Encapsulite filter type you can
} confidently recommend for use with Kodak SO163 EM Film, or
} would you prefer something else? If so, what? The darkroom will
} be 3x3metres, and ideally I want a diffuse, general room lighting
} that will serve for both EM film loading and processing and bromide
} paper printing.
}
} Chris
} ==========================================
} Dr. Chris Jeffree
} University of Edinburgh
} BIOSEM - Biological Sciences Electron Microscope Facility
} Institute of Cell and Molecular Biology
} Waddington Building
} King's Buildings, Mayfield Road
} EDINBURGH, EH9 3JN, Scotland, UK
} Tel. #44 (0) 131 650 5554
} FAX. #44 (0) 131 650 6563
} Mobile 07710 585 401
} email c.jeffree-at-ed.ac.uk
} =========================================
}
}
}


From daemon Wed Jun 19 19:13:33 2002



From: rszilard-at-sickkids.ca (rachel szilard)
Date: Wed, 19 Jun 2002 20:05:42 -0400
Subject: LM - uncooperative antibodies & immunofluorescence

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I am fairly new to staining tissues (mouse) and I am having trouble
detecting my protein of interest by indirect immunofluorescence. I have
raised antibodies (rabbit, guinea pig) to the cytoplasmic domain of a
neuronal transmembrane protein. These antibodies work well on
Westerns and in immunofluorescence of transfected COS cells (fixed
with 4% PFA in PBS, permeabilised with PBS + 0.1% Triton) However I
see little or nothing above background on sections of brain or retina.
The tissues are fixed overnight in 4% PFA prior to sectioning,
rehydrated in PBS and are permeabilised using 0.1% Triton before
blocking (PBS + 10% serum) and antibody addition. Antibodies to other
proteins, with various subcellular distributions, work well under these
conditions. I have tried post-fixing in 4% PFA without improvement. I
have not yet tried unfixed sections. If anyone has suggestions for
alternative fixation/permeabilisation/detection methods, or can direct me
to a good reference I would be very appreciative.


Thanks in advance,

rachel



Rachel K. Szilard, Ph.D.

Program in Developmental Biology
11-130 Elm Wing
Hospital for Sick Children
555 University Avenue
Toronto, Ontario
Canada
M5G 1X8

(416) 813 4997 phone
(416) 813 4931 fax
rszilard-at-sickkids.ca




From daemon Wed Jun 19 19:56:41 2002



From: Josh Kahn :      kahnj-at-biology.queensu.ca
Date: Wed, 19 Jun 2002 20:50:11 -0400 (EDT)
Subject: CLSM - Intermediate filament immunofluorescence of plant tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi listers,

I'm looking for protocols for immunostaining intact (not protoplasts)
plant tissue intermediate filaments (using anti IFA, ME 101, or anything
else that will light the buggers up! Has anyone used anti-keratin
primaries?). Ideally for pea root nodules but anything else will help.

Thanks!



From daemon Wed Jun 19 22:08:03 2002



From: Dmrelion-at-aol.com
Date: Wed, 19 Jun 2002 23:00:09 EDT
Subject: immersion oil removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear list members,

One of my associates has a large number of thin sections of limestone that he
wants to look at with cathodoluminescence (CL). The problem is that the thin
sections have been coated already with immersion oil for other microscopy
observations and this may create problems with the vacuum and electron beam
in the CL apparatus. Does anyone have suggestions on the best/easiest way to
remove the immersion oil? The oil is identified as "Immersion oil for
Microscopy, Cargille Type A, Formula Code 1248" and is said to be made up of
synthetic hydrocarbons and natural petroleum derivatives.

He will appreciate any and all suggestions.

Thank you.

Don Marshall

RELION Industries
PO Box 12
Bedford, MA 01730

781-275-4695 (Phone)
781-271-0252 (FAX)

dmrelion-at-aol.com

cathodoluminescence, mass spectroscopy, electron beam technology

http://www.excitingelectrons.com


From daemon Wed Jun 19 23:40:09 2002



From: Vitaly Feingold :      vitalylazar-at-worldnet.att.net
Date: Thu, 20 Jun 2002 00:05:51 -0400
Subject: Philips EM420 IGP Power Supply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Krzysztof, what's wrong with the IGP power supply? It is a very simple
circuit that is much easier and cheaper to repair than to ship overseas.
Components are available everywhere. Contact me off list with the detailed
description of the problem.

Vitaly Feingold
Scientific Instruments and Applications
2773 Heath Lane, Duluth GA 30096
(770)232-7785 ph.
(770)232-1791 fax
(678)467-0012 mobile

This message is made of 100% recycled electrons.
----- Original Message -----
} From: Krzysztof Herman {kherman-at-labsoft.com.pl}
To: MSA {Microscopy-at-sparc5.microscopy.com}
Sent: Wednesday, June 19, 2002 4:26 PM


Hello List

Does somebody have the "IGP power supply" (or maybe some of it's components)
to sell cheap
or even donate to Czech microscopists ??

(I don't count on many offers ;-)) ... but to ask is allways worthwise..

regards

Krzysztof Herman
Labsoft, ul.Bażancia 45A
02-892 Warszawa, Poland
tel/fx: (+48 22)6449753, 6449750
mobile: (+48 601)307456








From daemon Thu Jun 20 07:05:06 2002



From: JPMcCullough-at-fsl.gov.ie
Date: Thu, 20 Jun 2002 12:56:39 +0100
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Unsubscribe, please.



"The information transmitted is intended only for the person or entity to which it
is addressed and may contain confidential and/or privileged material. Any review,
retransmission, dissemination or other use of, or taking of any action in reliance
upon, this information by persons or entities other than the intended recipient is
prohibited. If you received this in error, please contact the sender and delete the
material from any computer. It is the policy of the Forensic Science Laboratory to
disallow the sending of offensive material and should you consider that the material
contained in this message is offensive you should contact the sender immediately and
also mailminder-at-fsl.gov.ie".


From daemon Thu Jun 20 07:58:52 2002



From: michael shaffer :      rarewolf-at-roadrunner.nf.net
Date: Thu, 20 Jun 2002 10:22:07 -0230
Subject: RE: immersion oil removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Don Marshall writes ...

} One of my associates has a large number of thin sections of
} limestone that he wants to look at with cathodoluminescence
} (CL). The problem is that the thin sections have been coated
} already with immersion oil ...
} ... Does anyone have suggestions on the
} best/easiest way to remove the immersion oil? ...

Ethanol should work ... but I generally cut the difficult stuff with a
little acetone (in a hood) first, then an immediate immersion in ETOH.

cheerios ... shAf :o)
Avalon Peninsula, Newfoundland
www.micro-investigations.com



From daemon Thu Jun 20 07:58:58 2002



From: Scott Whittaker :      Whittaker.scott-at-nmnh.si.edu
Date: Thu, 20 Jun 2002 08:52:49 -0400
Subject: Sputter target suppliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I only have two sources, Pella and Refining Systems. Who else is there??
Thanks

Scott Whittaker
SEM Lab Manager
Smithsonian Institution
PO Box 37012 MRC104
National Museum of Natural History
Washington DC 20013-7012
202-357-1651



From daemon Thu Jun 20 08:34:05 2002



From: Walck, Scott D. :      walck-at-ppg.com
Date: Thu, 20 Jun 2002 09:26:12 -0400
Subject: Jet polishing recipe for stainless steel 316LN

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have a couple of recipes from the Philips monographs# 5 book (Thompson-Russell and Edington).

1) 10% Perchloric 90% acetic, 12v, 0C, jet polish
for austenitic SS:
2) 10% Perchloric, 20% glycerol, 70% ethanol, 65V electropolish
3)10% nitric, 90% water 50V, jet polish
4) 60% orthophosphoric, 40% sulphuric, 100C, no voltage specified, Jet (this works well for electropolishing too.)

-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8515 (fax)



-----Original Message-----
} From: Yan Xin [mailto:xin-at-magnet.fsu.edu]
Sent: Wednesday, June 19, 2002 2:22 PM
To: Microscopy-at-sparc5.microscopy.com


Hi,

I need to make TEM samples from the bulk stainless steel.
I would appreciate very much if anyone could give me the recipe to make TEM
samples of stainless steel 316LN using jet polishing.

Thanks

Regards
Yan xin
=======================================
Yan Xin (Ph.D)
Magnet Science & Technology
National High Magnetic Field Laboratory
Florida State University
1800 E. Paul Dirac Drive
Tallahassee, FL 32310
Tel: (850) 644 1529
Fax: (850) 644 0867
========================================





From daemon Thu Jun 20 10:21:57 2002



From: jtwilley-at-sprynet.com
Date: Thu, 20 Jun 2002 11:08:44 -0400
Subject: Re: immersion oil removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Don,

I have had good success using a very light petroleum distillate such as hexane
or "petroleum ether". This has very low surface tension (penetrates porosity
well) and a high evaporation rate so that it may be repeated several times in
quick succession. I would use a tissue to remove the excess immersion oil
first, followed by immersion in a minimal volume of the solvent - say 5ml for
a standard slide. Drain, allow to dry for a minute and then repeat with a
fresh 5ml volume. I have never seen observable residues in the SEM even from
structures such as clay stylolites inside biocalcarinites. I haven't tried CL
however, after this.

John Twilley
Conservation Scientist

On Wed, 19 Jun 2002 23:00:09 EDT "Dmrelion-at-aol.com"-at-sparc5.microscopy.com
wrote:

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Dear list members,

One of my associates has a large number of thin sections of limestone that he

wants to look at with cathodoluminescence (CL). The problem is that the thin
sections have been coated already with immersion oil for other microscopy
observations and this may create problems with the vacuum and electron beam
in the CL apparatus. Does anyone have suggestions on the best/easiest way to
remove the immersion oil? The oil is identified as "Immersion oil for
Microscopy, Cargille Type A, Formula Code 1248" and is said to be made up of
synthetic hydrocarbons and natural petroleum derivatives.

He will appreciate any and all suggestions.

Thank you.

Don Marshall

RELION Industries
PO Box 12
Bedford, MA 01730

781-275-4695 (Phone)
781-271-0252 (FAX)

dmrelion-at-aol.com

cathodoluminescence, mass spectroscopy, electron beam technology

http://www.excitingelectrons.com




From daemon Thu Jun 20 12:45:09 2002



From: Anthony J. Garratt-Reed :      tonygr-at-mit.edu
Date: Thu, 20 Jun 2002 13:38:06 -0400
Subject: RE: Jet polishing recipe for stainless steel 316LN

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The recipes Scott gives would undoubtedly work, but there are significant
hazards, especially associated with Perchloric Acid. I would strongly urge
new users to discuss the hazards of this material with their Health and
Safety people before beginning work.

These hazards are not hypothetical. Several years ago we had a user
working with 20% Perchloric in Methanol (thankfully at -35C) in the
jet-polisher. For whatever reason (we have never found out) the solution
caught fire. Thankfully a clear-headed staff member was nearby and added
copious amounts of water, but the student was panicked, and frozen into
inaction. At higher temperatures this particular mixture is explosive!!!

We threw the jet polisher away and have not replaced it. All steel samples
in our lab now are made by ion-milling.

Tony

At 09:26 AM 6/20/2002 -0400, Walck, Scott D. wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6478
**




From daemon Thu Jun 20 13:27:36 2002



From: jerzy.gazda-at-amd.com
Date: Thu, 20 Jun 2002 13:20:18 -0500
Subject: Jet polishing recipe for stainless steel 316LN

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Yan Xin,
working with perchloric acid can be a dangerous business, be careful not to let any salts accumulate on outside of your solution bottles, cool the stuff while mixing, and do not store used solutions for extended period of time. Salts of perchloric acid can be explosive or spontaneously combust. Also in USA OSHA requires special hoods with water cleanup for work with perchloric acid.

Alternatively, I have a recipe from a guide book published years ago by Bernie Kestel at ANL. The reference for this publication is ANL-80-120 revised edition from 1986.

Using jet electro polisher (single polisher was used to acquire this data):

solution of:
20% H2SO4 + 80% Methanol at 20°C with medium stir rate, 20V, 320-400 mA gives removal rate of ~300nm / sec when fresh or 200nm when re-used.

Personally, I like to mix these components while holding my container in a cooling bath of methanol and dry ice (solid CO2), they heat up and could be flammable when combined too fast. I used this solution extensively with SS316 and vanadium alloys (irradiated and annealed materials) and found that mixing the two types of specimens did not cause any problems, as mixed solution has about one month shelf life due to its hygroscopic nature. Cooling the solution below 0°C allows better control of removal rates.


Have fun and be careful to work in a vented hood.

Jerzy
******************************************************
Jerzy Gazda, Ph.D. Advanced Micro Devices
Supervising Engineer 5204 E. Ben White Blvd. - MS 613
PCAL - AIM Section Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453 FAX: (512) 602-7470
jerzy.gazda-at-amd.com
******************************************************







-----Original Message-----
} From: Walck, Scott D. [mailto:walck-at-ppg.com]
Sent: Thursday, June 20, 2002 8:26 AM
To: 'Yan Xin'
Cc: Microscopy (E-mail)


I have a couple of recipes from the Philips monographs# 5 book (Thompson-Russell and Edington).

1) 10% Perchloric 90% acetic, 12v, 0C, jet polish
for austenitic SS:
2) 10% Perchloric, 20% glycerol, 70% ethanol, 65V electropolish
3)10% nitric, 90% water 50V, jet polish
4) 60% orthophosphoric, 40% sulphuric, 100C, no voltage specified, Jet (this works well for electropolishing too.)

-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8515 (fax)



-----Original Message-----
} From: Yan Xin [mailto:xin-at-magnet.fsu.edu]
Sent: Wednesday, June 19, 2002 2:22 PM
To: Microscopy-at-sparc5.microscopy.com


Hi,

I need to make TEM samples from the bulk stainless steel.
I would appreciate very much if anyone could give me the recipe to make TEM
samples of stainless steel 316LN using jet polishing.

Thanks

Regards
Yan xin
=======================================
Yan Xin (Ph.D)
Magnet Science & Technology
National High Magnetic Field Laboratory
Florida State University
1800 E. Paul Dirac Drive
Tallahassee, FL 32310
Tel: (850) 644 1529
Fax: (850) 644 0867
========================================







From daemon Thu Jun 20 14:34:27 2002



From: Gillmeister, Russ :      RGillmeister-at-crt.xerox.com
Date: Thu, 20 Jun 2002 16:43:42 -0400
Subject: Immersion oil removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,
I know this is a little "off list", but does anyone out there have
any any strong opinions (oh boy, am I asking for it) about xylene vs
d-Limonene as a clearing solvent for paraffin processing? We
switched to limonene a little while ago, because of its lower
toxicity, but have had problems with adequate infiltration and
crunchy tissues, especially liver.
thanks in advance. You can respond on or off list,
Lee

--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From root Thu Jun 20 15:35:46 2002
Return-Path: {Microscopy-request-at-sparc5.microscopy.com}
Received: (from daemon-at-localhost)
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id PAA14150
for dist-Microscopy; Thu, 20 Jun 2002 15:07:11 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id PAA14145
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Thu, 20 Jun 2002 15:06:40 -0500 (CDT)
Received: from rwja.umdnj.edu ([130.219.4.100])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id PAA14134
for {microscopy-at-sparc5.microscopy.com} ; Thu, 20 Jun 2002 15:06:25 -0500 (CDT)
Received: from umdnj.edu ([130.219.170.159])
by rwja.umdnj.edu (8.9.3 (PHNE_24419)/8.9.3) with ESMTP id QAA26497
for {microscopy-at-sparc5.microscopy.com} ; Thu, 20 Jun 2002 16:01:42 -0400 (EDT)
Message-ID: {3D123548.40EA3AD-at-umdnj.edu}


Dear List:

I have a nice late 1950's vintage Wild M-11 light microscope. The
coarse focus mechanism is shot, the teeth on the two pieces that
interact are beyond repair. There don't seem to be any spare parts
available, even from Heerbrugg, Switzerland. I am looking for someone
with parts for this microscope or another M-11 with a different
unrepairable defect that I can cannibalize for parts.

Geoff
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************




From root Thu Jun 20 15:51:00 2002
Return-Path: {Microscopy-request-at-sparc5.microscopy.com}
Received: (from daemon-at-localhost)
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id PAA14204
for dist-Microscopy; Thu, 20 Jun 2002 15:14:56 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id PAA14198
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Thu, 20 Jun 2002 15:14:26 -0500 (CDT)
Received: from mailhub2.otago.ac.nz (mailhub2.otago.ac.nz [139.80.64.199])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id PAA14188
for {microscopy-at-sparc5.microscopy.com} ; Thu, 20 Jun 2002 15:14:13 -0500 (CDT)
Received: (from root-at-localhost)
by mailhub2.otago.ac.nz (8.10.1/8.10.1) id g5KK9Vd19024;
Fri, 21 Jun 2002 08:09:31 +1200
Received: from galadriel.otago.ac.nz (galadriel.otago.ac.nz [139.80.64.213])
by mailhub2.otago.ac.nz (8.10.1/8.10.1) with ESMTP id g5KK9UC18986;
Fri, 21 Jun 2002 08:09:30 +1200
Received: from [139.80.34.81] (ou034081.otago.ac.nz [139.80.34.81])
by galadriel.otago.ac.nz (8.11.2/8.11.2) with ESMTP id g5KK9RI07600;
Fri, 21 Jun 2002 08:09:28 +1200 (NZST)
Mime-Version: 1.0
X-Sender: st004104-at-brandywine.otago.ac.nz
Message-Id: {a0432040db937e57f8d7e-at-[139.80.34.81]}


Kia Ora Scott,

We send our sputter coater targets for recoating to

Testbourne Ltd
Unit 12/13, Hassocks Wood,
Stroudley Road
Basingstoke
Hampshire,
England
RG24 8UQ

Phone ++44 (0) 1256 467055
Fax ++44 (0) 1256 842929

email: info-at-testbourne.com


Good turn around time (and we are on the other side of the world),
good price, they give us a refund on any metal scraps left on the
target, and they send us a christmas card at Christmas...... what
more could you ask for.


Allan




At 8:52 AM -0400 20/6/02, Scott Whittaker wrote:
} I only have two sources, Pella and Refining Systems. Who else is there??
} Thanks
}
} Scott Whittaker
} SEM Lab Manager
} Smithsonian Institution
} PO Box 37012 MRC104
} National Museum of Natural History
} Washington DC 20013-7012
} 202-357-1651
--
-------------------------------------------------
Allan Mitchell
Technical Manager
Otago Centre for Electron Microscopy
C/-Department of Anatomy and Structural Biology
School of Medical Sciences
P.O. Box 913
Dunedin
New Zealand

Phone (03) 479 5642 or 479 7301
Fax (03) 479 7254

Unit: http://www.otago.ac.nz/anatomy/emunit/
Department: http://anatomy.otago.ac.nz/



"Life is a gift, don't waste it"


From root Thu Jun 20 16:21:08 2002
Return-Path: {Microscopy-request-at-sparc5.microscopy.com}
Received: (from daemon-at-localhost)
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id PAA14372
for dist-Microscopy; Thu, 20 Jun 2002 15:49:22 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id PAA14368
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Thu, 20 Jun 2002 15:48:52 -0500 (CDT)
Received: from umbriel.eastgw.xerox.com (umbriel.xerox.com [208.140.33.26])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id PAA14361
for {Microscopy-at-sparc5.microscopy.com} ; Thu, 20 Jun 2002 15:48:37 -0500 (CDT)
Received: from crte147.wc.eso.mc.xerox.com (crte147.wc.eso.mc.xerox.com [13.1.158.3])
by umbriel.eastgw.xerox.com (8.9.3/8.9.3) with ESMTP id QAA22466;
Thu, 20 Jun 2002 16:43:22 -0400 (EDT)
Received: by crte147.wc.eso.mc.xerox.com with Internet Mail Service (5.5.2653.19)
id {M6Y0ZQKF} ; Thu, 20 Jun 2002 16:43:47 -0400
Message-ID: {BF85F885643CD21188230008C75629350476176A-at-crte128-2.wrc.xerox.com}


Don, I'm digging way back in the gray matter for this one. I believe it was
Ladd Ind. who sold a reflux unit specifically for the purpose of washing TEM
replicas in the days before SEM. It consisted of a water cooled probe which
supported the grid on a fine wire mesh. This probe was inserted into the
reflux unit where the heated solvent vapors would continually wash the grid.
You could probably build a similar setup and use Hexane to wash out the oil.


Russ Gillmeister
Microscopy
Xerox Corp.
Webster, NY 14580




From daemon Thu Jun 20 17:12:08 2002



From: Michael Calabrese :      mcalabrese-at-rwsc.com
Date: Thu, 20 Jun 2002 16:08:15 -0700
Subject: RE: Jet polishing recipe for stainless steel 316LN

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Energy Beam Sciences.

Michael R. Nesta
General Manager
Energy Beam Sciences, Inc.
Agawam, MA USA
Tel: (800) 992-9037
www.ebsciences.com
"Adding Brilliance to Your Vision"



-----Original Message-----
} From: Scott Whittaker [mailto:Whittaker.scott-at-nmnh.si.edu]
Sent: Thursday, June 20, 2002 8:53 AM
To: microscopy-at-microscopy.com


I only have two sources, Pella and Refining Systems. Who else is there??
Thanks

Scott Whittaker
SEM Lab Manager
Smithsonian Institution
PO Box 37012 MRC104
National Museum of Natural History
Washington DC 20013-7012
202-357-1651





From root Thu Jun 20 18:35:30 2002
Return-Path: {Microscopy-request-at-sparc5.microscopy.com}
Received: (from daemon-at-localhost)
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id SAA15640
for dist-Microscopy; Thu, 20 Jun 2002 18:13:09 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id SAA15637
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Thu, 20 Jun 2002 18:12:38 -0500 (CDT)
Received: from photon.rwsc.com ([130.50.23.2])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id SAA15630
for {Microscopy-at-sparc5.microscopy.com} ; Thu, 20 Jun 2002 18:12:26 -0500 (CDT)
Received: (from uucp-at-localhost)
by photon.rwsc.com (8.11.6+Sun/8.11.6) id g5KN62n25081
for {Microscopy-at-sparc5.microscopy.com} ; Thu, 20 Jun 2002 16:06:02 -0700 (PDT)
Received: from dns.risc.rockwell.com(130.50.4.15) by photon.rwsc.com via csmap (V6.0)
id srcAAAmTaa_W; Thu, 20 Jun 02 16:06:01 -0700
Received: from rsctonotes06.rsc.rockwell.com (rsctonotes06.risc.rockwell.com [130.50.4.76])
by dns.rwsc.com (8.11.6+Sun/8.10.2) with ESMTP id g5KN7hI12862
for {Microscopy-at-sparc5.microscopy.com} ; Thu, 20 Jun 2002 16:07:43 -0700 (PDT)



I have used the following solutions for jet polishing(Tenupol) various
stainless steels and Ni based alloys:
a) 125ml methanol, 75ml butyl alcohol, 20ml perchloric acid -at-
-35C/16V

b) 150ml methanol, 50ml nitric acid -at- -40C/15V

They both work well. There are dangers if you heat these solutuions. Cool
the methanol before adding the acid. Keep them cool when using and do not
store. Dilute them with water when transfering to a disposal container.
(Don't catch them on fire!)

Mike Calabrese
mcalabrese-at-rwsc.com



From daemon Thu Jun 20 18:54:52 2002



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Thu, 20 Jun 2002 21:32:44 -0500
Subject: Removal of Cargille Type A, Formula Code 1248

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Try:

Refining Systems, Inc.
Las Vegas, NV
702-368-0579.


gg


} At 8:52 AM -0400 20/6/02, Scott Whittaker wrote:
} } I only have two sources, Pella and Refining Systems. Who else is there??
} } Thanks
} }
} } Scott Whittaker
} } SEM Lab Manager
} } Smithsonian Institution
} } PO Box 37012 MRC104
} } National Museum of Natural History
} } Washington DC 20013-7012
} } 202-357-1651



From root Thu Jun 20 21:05:36 2002
Return-Path: {Microscopy-request-at-sparc5.microscopy.com}
Received: (from daemon-at-localhost)
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id UAA16073
for dist-Microscopy; Thu, 20 Jun 2002 20:38:02 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id UAA16070
for "MicroscopyFilteredEmail1-at-msa.microscopy.com"; Thu, 20 Jun 2002 20:37:32 -0500 (CDT)
Received: from diskless5.axs2000.net (mail.axs2000.net [209.120.196.43])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id UAA16063
for {Microscopy-at-sparc5.Microscopy.com} ; Thu, 20 Jun 2002 20:37:20 -0500 (CDT)
Received: from [64.80.72.149] (host-072-149.paetec.axs2000.net [64.80.72.149])
by diskless5.axs2000.net (8.11.2/8.11.2) with SMTP id g5L1TFc17990
for {Microscopy-at-MSA.Microscopy.com} ; Thu, 20 Jun 2002 21:29:16 -0400
Message-Id: {200206210129.g5L1TFc17990-at-diskless5.axs2000.net}
To: MICROSCOPY BB {Microscopy-at-sparc5.microscopy.com}


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Don Marshall wrote:
=============================================================
One of my associates has a large number of thin sections of limestone that
he
wants to look at with cathodoluminescence (CL). The problem is that the thin
sections have been coated already with immersion oil for other microscopy
observations and this may create problems with the vacuum and electron beam
in the CL apparatus. Does anyone have suggestions on the best/easiest way to
remove the immersion oil? The oil is identified as "Immersion oil for
Microscopy, Cargille Type A, Formula Code 1248" and is said to be made up of
synthetic hydrocarbons and natural petroleum derivatives.

He will appreciate any and all suggestions.
=============================================================
The manufacturer of Cargille fluids told me today that the "recommended"
solvents would be heptane, tolune, xylene, or ethyl ether. Interestingly,
they said acetone would not do as good of a job as these other solvents.

Disclaimer: SPI Supplies distributes the full range of fluids from Cargille
for light microscopy applications.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================




From daemon Thu Jun 20 22:57:59 2002



From: Frank Eugene Jones :      jonesfe-at-darkwing.uoregon.edu
Date: Thu, 20 Jun 2002 20:46:41 -0700
Subject: Re: Sputter target suppliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Scott,

Try the Kurt J. Lesker company. I see on their web site
(www.kurtlesker.com) that they are having a sale on targets.

Frank
}
----------------------------------------------------------------------
--
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
----------------------------------------------------------------------
-.
}
}
} I only have two sources, Pella and Refining Systems. Who else is
there??
} Thanks
}
} Scott Whittaker
} SEM Lab Manager
} Smithsonian Institution
} PO Box 37012 MRC104
} National Museum of Natural History
} Washington DC 20013-7012
} 202-357-1651
}
}
*******************************************************
Frank E. Jones Physics Graduate Student
Lonergan Lab (Chemistry) University of Oregon
Office: 103a Klamath Phone: (541)346-2900
or 346-0977
email: jonesfe-at-darkwing.uoregon.edu
homepage: darkwing.uoregon.edu/~jonesfe


From daemon Fri Jun 21 01:02:20 2002



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Fri, 21 Jun 2002 07:52:28 +0200
Subject: RE: immersion oil removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Don

I am passing on a bit of advice a service technician give me to clean
microscope slides from immersion oil. This may work for your sample as
well. Normally I just rinse slides with it. Since your sample is more
robust the following may work. :



Place the with the exposed side down for approx 2 minutes in a mixture of
10 % Acetone
20 % Chloroform
70 % Absolute Ethanol

You can use a hair drier to speed up the evaporation




Mr S. H. Coetzee
Electron Microscope Unit
University of Botswana
Private Bag 0022
Gabarone
Botswana

Phone : +267 355 2426
Fax : +267 585 097
e-mail : coetzees-at-mopipi.ub.bw


} -----Original Message-----
} From: "Dmrelion-at-aol.com"-at-sparc5.microscopy.com
} [mailto:"Dmrelion-at-aol.com"-at-sparc5.microscopy.com]
} Sent: Thursday, June 20, 2002 5:00 AM
} To: microscopy-at-microscopy.com
} Subject: immersion oil removal
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} --------------------------------------------------------------
} ---------.
}
}
} Dear list members,
}
} One of my associates has a large number of thin sections of
} limestone that he
} wants to look at with cathodoluminescence (CL). The problem
} is that the thin
} sections have been coated already with immersion oil for
} other microscopy
} observations and this may create problems with the vacuum and
} electron beam
} in the CL apparatus. Does anyone have suggestions on the
} best/easiest way to
} remove the immersion oil? The oil is identified as
} "Immersion oil for
} Microscopy, Cargille Type A, Formula Code 1248" and is said
} to be made up of
} synthetic hydrocarbons and natural petroleum derivatives.
}
} He will appreciate any and all suggestions.
}
} Thank you.
}
} Don Marshall
}
} RELION Industries
} PO Box 12
} Bedford, MA 01730
}
} 781-275-4695 (Phone)
} 781-271-0252 (FAX)
}
} dmrelion-at-aol.com
}
} cathodoluminescence, mass spectroscopy, electron beam technology
}
} http://www.excitingelectrons.com
}


From daemon Fri Jun 21 01:06:54 2002



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Fri, 21 Jun 2002 07:58:59 +0200
Subject: RE: Sputter target suppliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Scott

I do not know how limited your fund are. But in the past we just buy 24
Carat gold in sheet form from the reserve bank. There is lot of additional
effort in getting permits etc. Then if you have a polaron which uses a ring
rather than a circle like in the SPI coater you still have to cut it
yourself which will kneed help from a workshop. The end result was that we
were able to get double the sputter target thickness for half the price. A
chat to a local jeweller will help.






Mr S. H. Coetzee
Electron Microscope Unit
University of Botswana
Private Bag 0022
Gabarone
Botswana

Phone : +267 355 2426
Fax : +267 585 097
e-mail : coetzees-at-mopipi.ub.bw


} -----Original Message-----
} From: Scott Whittaker [mailto:Whittaker.scott-at-nmnh.si.edu]
} Sent: Thursday, June 20, 2002 2:53 PM
} To: microscopy-at-microscopy.com
} Subject: Sputter target suppliers
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} --------------------------------------------------------------
} ---------.
}
}
} I only have two sources, Pella and Refining Systems. Who else
} is there??
} Thanks
}
} Scott Whittaker
} SEM Lab Manager
} Smithsonian Institution
} PO Box 37012 MRC104
} National Museum of Natural History
} Washington DC 20013-7012
} 202-357-1651
}
}


From daemon Fri Jun 21 07:50:01 2002



From: John Hoffpauir :      John.Hoffpauir-at-mail.tju.edu
Date: Fri, 21 Jun 2002 08:42:34 -0400
Subject: Re: Sputter Coater Targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


hello all, does anyone out there know how to make up ammonium carbonate buffer
to pH 7.0.
thanks
john


From daemon Fri Jun 21 12:11:22 2002



From: Greg Barclay :      gbarclay-at-trinidad.net
Date: Thu, 20 Jun 2002 12:59:13 -0500
Subject: Olympus DP12 digital camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,
I am considering buying an Olympus DP12 digital camera for light microscopy use and I am looking for advice on how this
camera compares with other cameras on the market from Nikon, Kodak, Zeiss etc.

I want something that is a good value now in the face of rising megapixels and falling prices. Is its 3.34 MP chip enough for
high quality images?

I would appreciate your comments because I have no experience with digital photography.

Greg




From daemon Fri Jun 21 14:57:10 2002



From: Andreas Taubert :      taubert-at-seas.upenn.edu
Date: Fri, 21 Jun 2002 15:48:52 -0700
Subject: IR microscope manufacturers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,

we are looking into buying an IR microscope and I am currently trying to
assemble a list of vendors. I am aware of Thermo Nicolet, Omicron, and
Perkin Elmer. Any other sources ? We are looking for an intrument that can
do microspectroscopy, i.e., spatially resolved IR spectroscopy on ~ 1 micron
(or better, of course) length scale.

Thanks in advance
Andreas

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Andreas Taubert
Dept. of Materials Science & Engineering
3231 Walnut St.
University of Pennsylvania
Philadelphia PA 19104-6272
Tel.: 215 573 1179
Fax: 215 573 2128
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~



From daemon Fri Jun 21 15:24:33 2002



From: Andreas Taubert :      taubert-at-seas.upenn.edu
Date: Fri, 21 Jun 2002 16:17:49 -0700
Subject: IR microscope manufacturers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,

we are looking into buying an IR microscope and I am currently trying to
assemble a list of vendors. I am aware of Thermo Nicolet, Omicron, and
Perkin Elmer. Any other sources ? We are looking for an intrument that can
do microspectroscopy, i.e., spatially resolved IR spectroscopy on ~ 1 micron
(or better, of course) length scale.

Thanks in advance
Andreas

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Andreas Taubert
Dept. of Materials Science & Engineering
3231 Walnut St.
University of Pennsylvania
Philadelphia PA 19104-6272
Tel.: 215 573 1179
Fax: 215 573 2128
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~



From daemon Fri Jun 21 15:31:28 2002



From: =?iso-8859-1?Q?Jos=E9?= Reinaldo Guerrero :      guerrero-at-ivic.ve
Date: Fri, 31 May 2002 18:05:48 -0400
Subject: subscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sir

Could you subscribe me at the microscopy club?

Thanks

José Reinaldo Guerrero





From daemon Fri Jun 21 17:13:15 2002



From: Mary K. O'Connell :      oconne1l-at-leland.stanford.edu
Date: Fri, 21 Jun 2002 15:05:11 -0700
Subject: Measurement of Collagen Fiber Orientation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopy Members:

I am struggling to measure the orientation of collagen fibers in the layers
of a blood vessel. In healthy aorta, I expect to see a semi-organized
arrangement of predominantly helically-oriented fibers, as well as a few
longitudinally-oriented fibers.

Thus far, have investigated polarized light microscopy (PLM) and confocal
laser scanning microscopy (CLSM). Initial efforts with CLSM revealed poor
collagen visibility (trying to improve staining technique) and difficulties
in imaging through the thickness of the vessel (0.2mm). PLM equipment was
limited and was difficult to correlate orientation from one layer to the
next.

I would greatly appreciate any input on possible techniques or resources
for measuring fiber orientation. Thank You!

Mary


***************************************************
Mary K.O'Connell
Cardiovascular Biomechanics Research Lab
MSLS: Room P224 - Surgery
1201 Welch Road
Stanford, CA 94305
(650) 723-1695
(650) 498-6262 Fax
E-mail: oconne1l-at-leland.stanford.edu
(Note: O'Connell in the E-mail address is spelled with a numeric "1" for
the first l, and an alphabetic "l" for the second.)
*******************************************************


From daemon Fri Jun 21 17:18:07 2002



From: Mary K. O'Connell :      oconne1l-at-leland.stanford.edu
Date: Fri, 21 Jun 2002 15:12:09 -0700
Subject: Measurement of Collagen Fiber Orientation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopy Members:

I am struggling to measure the orientation of collagen fibers in the layers
of a blood vessel. In healthy aorta, I expect to see a semi-organized
arrangement of predominantly helically-oriented fibers, as well as a few
longitudinally-oriented fibers.

Thus far, have investigated polarized light microscopy (PLM) and confocal
laser scanning microscopy (CLSM). Initial efforts with CLSM revealed poor
collagen visibility (trying to improve staining technique) and difficulties
in imaging through the thickness of the vessel (0.2mm). PLM equipment was
limited and was difficult to correlate orientation from one layer to the
next.

I would greatly appreciate any input on possible techniques or resources
for measuring fiber orientation. Thank You!

Mary


***************************************************
Mary K.O'Connell
Cardiovascular Biomechanics Research Lab
MSLS: Room P224 - Surgery
1201 Welch Road
Stanford, CA 94305
(650) 723-1695
(650) 498-6262 Fax
E-mail: oconne1l-at-leland.stanford.edu
(Note: O'Connell in the E-mail address is spelled with a numeric "1" for
the first l, and an alphabetic "l" for the second.)
*******************************************************


From daemon Fri Jun 21 18:03:28 2002



From: Bill & Sue Tivol :      wtivol-at-earthlink.net
Date: Fri, 21 Jun 2002 15:55:13 -0400
Subject: Re: Ammonium carbonate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


on 6/21/02 8:42 AM, John Hoffpauir at John.Hoffpauir-at-mail.tju.edu wrote:
}
} hello all, does anyone out there know how to make up ammonium carbonate buffer
} to pH 7.0.
} thanks
} john
}
Dear John,
Ammonium bicarbonate (NH4HCO3) is pretty close to pH 7.0. If it is
below that pH, just add (NH4)2CO3; if it is above that pH, just bubble CO2
through the solution.
Yours,
Bill Tivol



From daemon Fri Jun 21 20:14:58 2002



From: Greg Barclay :      gbarclay-at-trinidad.net
Date: Thu, 20 Jun 2002 21:03:55 -0500
Subject: Re: LM: need help fixing + embedding sugarcane

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Gabrielle,

Young sugarcane internodes (cross sections abt 1-1.5 cm diam) and stem apices (longitudinal sections) fix fine in an
FAA-TBA-Ethanol series we have been using at the University of the West Indies in Trinidad, forever and ever. If you
send me some more details of your methods I will ask my technician next week, and I can send you our fixation
schedule, which is as old as the hills.

Greg


} Does anyone have any experience with fixing and embedding
} sugarcane stem tissue for in situ hybridization? So far we have
} tried a number of protocols - including traditional fixation in 4%
} paraformaldehyde, as well as prefixing and then freezing in
} isopentane and cutting frozen sections - but to no avail. With
} traditional chemical fixation (a protocol which worked beautifully for
} one of our colleagues working on grape berries), we have to fix for 5-
} 7 days just to get adequate fixation, and even then parts of the
} older internode tissue don't fix properly - this makes sectioning very
} tricky. With cutting prefixed frozen tissue, our sections break up
} very easily, and even when we do get sections, once thawed they
} go into some sort of shock and develop thick black cell borders
} (even in the xylem).
}
} If anyone has any ideas, please let me know!
}
} Thanks
} Gabrielle
}
}
}
} Gabrielle Turner
} Institute for Plant Biotechnology
} University of Stellenbosch
} 7600
}
} Cell: 083 324 7453
}





From daemon Fri Jun 21 20:32:29 2002



From: Greg Barclay :      gbarclay-at-trinidad.net
Date: Thu, 20 Jun 2002 21:23:31 -0500
Subject: Re: Need help fixing and embedding Sugarcane

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello Gabrielle,

Young sugarcane internodes (cross sections abt 1-1.5 cm diam) and whole stem apices section great after fixation in an
FAA-TBA-Ethanol series we have been using at the University of the West Indies in Trinidad forever.

Send me some more details of your methods and I will ask my technician next week to check your schedule. I can send
you our schedule which is as old as the hills.

Greg





From daemon Fri Jun 21 20:51:05 2002



From: Greg Barclay :      gbarclay-at-trinidad.net
Date: Thu, 20 Jun 2002 21:41:48 -0500
Subject: Reply to Need help fixing and embedding Sugarcane

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello Gabrielle,
}
} Young sugarcane internodes (cross sections abt 1-1.5 cm diam) and whole stem apices section great in an
} FAA-TBA-Ethanol series we have been using at the University of the West Indies in Trinidad.
}
} Send me some more details of your methods and I will ask my technician next week to look at them. I can send you our
schedule, which is as old as the hills.
}
} Greg




From daemon Sat Jun 22 19:38:52 2002



From: DrJohnRuss-at-aol.com
Date: Sun, 23 Jun 2002 07:29:51 EDT
Subject: Re: Measurement of Collagen Fiber Orientation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mary,

I was using my BioRad MRC1000 CLSM) to look at sections of some cells grown
in collagen matrix and stained with H&E. I was amazed that the collagen
really lit up as did the cells but not the nuclei. These were cross section
of rat aorta and I was looking at the endothelial cells growing out to form
accessory capillaries? The sections were about 30-50 micrometers thick.

As far a orientation, Fovea Pro 2.0 sold by Reindeer Games can do that with
a digital image; I'm not sure about a stack of images. Check with Dr. John
Russ, he can advise you better; he'll probably answer your email anyway.
Usual disclaimer - no financial interest, etc. just a user.

Damian

Damian Neuberger, Ph.D
Senior Research Scientist
Baxter Healthcare Corp.

----- Original Message -----
} From: "Mary K. O'Connell" {oconne1l-at-leland.stanford.edu}
To: {Microscopy-at-sparc5.microscopy.com}
Sent: Friday, June 21, 2002 5:12 PM



In a message dated 6/22/02 8:51:23 PM, neuberger1234-at-attbi.com writes:

} As far a orientation, Fovea Pro 2.0 sold by Reindeer Games can do that
} with a digital image; I'm not sure about a stack of images. Check with Dr.
} John Russ, he can advise you better; he'll probably answer your email anyway.

Well, I guess I'd better answer it now. I have no advice on sample prep or
getting the image in the first place, but if you have representative images
of the structure the procedure for determining orientation is pretty
straightforward, and yes, Fovea Pro (http://reindeergraphics.com) does
include that function. At each point in the image, the vector of brightness
gradient points perpendicular to the orientation of the collagen fiber.
Calculating that angle and assigning the value as a grey scale to the pixel
is the first step. If there are points in the image that are not collagen
fibers, a Boolean operation is then used to eliminate them. Finally, the
histogram of the grey scale (orientation) values is summed and plotted as a
rose plot to show the preferred orientation.

John Russ


From daemon Sun Jun 23 08:56:49 2002



From: Brendan J. Griffin :      bjg-at-cyllene.uwa.edu.au (by way of
Date: Sun, 23 Jun 2002 08:44:57 -0500
Subject: Lecturer Position: Centre for Microscopy & Microanalysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



LECTURER (REF: A26/20)
CENTRE FOR MICROSCOPY AND MICROANALYSIS
The Centre for Microscopy and Microanalysis (CMM) is the principal
node of the Western Australian Centre for Microscopy (WACM), a State
Government Centre of Excellence at the University of Western
Australia. WACM has partner facilities at Curtin, Murdoch and Edith
Cowan Universities. The CMM is also a core partner in the recently
established Nanostructural Analysis Network Organisation (NANO) and
we expect to install a high-resolution ion microprobe in 2003, as
part of the local NANO capability. CMM facilities are extensive, and
include digital optical and confocal microscopes, four TEMs and seven
SEMs, together with extensive specimen preparation equipment in a
newly refurbished site.

The appointee(s) will provide leadership and collaboration with users
in biological research involving optical, confocal and electron
microscopy techniques. In particular, the appointee will be expected
to initiate the development and enhancement of cryo-microscopy, x-ray
microanalysis and associated specimen preparation. Activities will
include the training and teaching of users of microscopical and
related microanalysis facilities in biological applications. It is
essential that the appointee has detailed knowledge in the
instrumental and application aspects of biological microscopy.
Experience with or an interest in relevant SIMS techniques would be
of benefit and should be noted in the application.

The position is expected to be formally shared with one or more of
the newly-formed UWA Schools of Plant Biology, Biomedical and
Chemical Sciences, Animal Biology or Anatomy and Human Biology.

The minimum qualification required is a PhD in Biological/Biomedical
Sciences or equivalent. The successful applicant must have strong
interpersonal skills and the ability to work with other staff within
the Centre and its users. Joint supervision of post-graduate students
is expected, in the relevant Schools. The position is tenurable.
Applicants with tertiary teaching experience should submit a teaching
portfolio as part of their application.

For further information and copies of the selection criteria please
contact the Director of the CMM, Associate Professor Brendan Griffin,
on telephone (08) 9380 2770 or email director-at-cmm.uwa.edu.au.
Details of the CMM may be found at: http://cmm.uwa.edu.au

SALARY RANGE: Lecturer Level B $53,592 - $63,642 p.a. (AUD)

CLOSING DATE: 28th June 2002

Late applications will continue to be received after the closing date
and may be considered until the position is filled. The University
reserves the right to not make an appointment to this position or to
fill the position at a different level.

Located adjacent to the picturesque banks of the Swan River, the
University of Western Australia offers an attractive benefits package
including generous superannuation, fares to Perth (if applicable) for
appointee and dependants along with a removals allowance, generous
leave provisions and a working environment that's the envy of many.
These and other benefits will be specified in the offer of employment.

APPLICATION DETAILS: Written applications quoting the reference
number, personal contact details, qualifications and experience,
along with contact detail of three referees should be sent to
Director, Human Resources, The University of Western Australia, 35
Stirling Highway, Crawley WA 6009 by the closing date.

Advertised internationally.

--




Brendan J. Griffin
Director & Assoc. Professor in Electron Microscopy
Centre for Microscopy and Microanalysis
The University of Western Australia
First floor, Physics Building,
35 Stirling Highway
CRAWLEY, WA, AUSTRALIA 6009
ph 61-8-9380-2739 fax 9380-1087
mobile 0409-104-096
bjg-at-cmm.uwa.edu.au
http://cmm.uwa.edu.au/

CRICOS Provider No. 00126G


From daemon Sun Jun 23 16:03:56 2002



From: Pauline C. Yu :      pcy-at-usc.edu
Date: Sun, 23 Jun 2002 13:55:44 -0700 (PDT)
Subject: light microscopy and RSIs--motorized stages?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,

I am currently experiencing possible symptoms of carpel tunnel syndrome
after having spent much of this summer at the helm of a microscope, and
this is in addition to the mild tendonitis I contracted from computer use
a few years ago. Does any one know about add-on motors for stages that
allow foot pedal control? I feel like this is a more viable solution than
buying a new scope with all the bells and whistles(besides the fact that
I'm only a graduate student and the budget decisions are not mine to
make!)
Of course any other suggestions from people with similar problems would be
greatly appreciated.
Thanks in advance,

Pauline Yu
pcy-at-usc.edu
Manahan Lab
http://www.usc.edu/manahanlab



From daemon Sun Jun 23 17:07:24 2002



From: msandusk-at-bucknell.edu (by way of MicroscopyListserver)
Date: Sun, 23 Jun 2002 16:56:39 -0500
Subject: Ask-A-Microscopist: immunostain & shrinkage question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (msandusk-at-bucknell.edu) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html#form on Sunday,
June 23, 2002 at 14:45:43
---------------------------------------------------------------------------

Email: msandusk-at-bucknell.edu
Name: Matt Sandusky

Organization: Bucknell Univeristy

Education: Undergraduate College

Location: Lewisburg, Pennsylvania, USA

Question: I'm working on an immunostain of 10 micron sections
of rat testis and I'm having trouble with shrinkage.
The seminiferous tubules are shrinking away from the
interstitial tissue. I'm currently working on perfusion
fixation using 4% paraformaldehyde and a .01M phosphate
buffer. Once fixed, I slice each testis in half and
embed in paraffin using 30min PBS, 50%EtOH for 1 hr,
70%EtOH 1hr, 90%EtOH 1hr, 3 of 100%EtOH 2hrs, 3 of
xylene for 1hr 30min, 2hr paraffin and 3hr paraffin (with vaccum).
Any advice anyone might be able to offer would be
greatly appreciated. Thanks.

---------------------------------------------------------------------------


From daemon Mon Jun 24 06:08:27 2002



From: Sampath, Srinidhi (CORP, GEITC) :      Srinidhi.Sampath-at-geind.ge.com
Date: Mon, 24 Jun 2002 16:20:56 +0530
Subject: 3D image analysis...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have an idea of how to do 3D image analysis for accurate
interparticle spacing on TEM images (projected images). Is there a
corresponding formula to:
lambda = d(particle)*f(matrix)/f(particle)
where lambda = interparticle spacing
d(particle) = dia of particle
f(matrix) = volume fraction of matrix
f(particle) = volume fraction of particle

Thanks,
Srinidhi
srinidhi.sampath-at-geind.ge.com


"THIS E-MAIL MESSAGE ALONG WITH ANY ATTACHMENTS IS INTENDED ONLY FOR THE
ADDRESSEE and may contain confidential and privileged information.
If the reader of this message is not the intended recipient,
you are notified that any dissemination, distribution or copy of this
communication is strictly Prohibited.
If you have received this message by error, please notify us
immediately, return the original mail to the sender and delete the
message from your system."



From daemon Mon Jun 24 07:37:24 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Mon, 24 Jun 2002 08:33:35 -0400
Subject: IR microscope manufacturers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Try Barnes also for thermal imaging. What wavelength do you need to work
at?

Peter Tomic
Anadigics, Inc.

-----Original Message-----
} From: Andreas Taubert [mailto:taubert-at-seas.upenn.edu]
Sent: Friday, June 21, 2002 6:49 PM
To: Microscopy List Server


Dear Listers,

we are looking into buying an IR microscope and I am currently trying to
assemble a list of vendors. I am aware of Thermo Nicolet, Omicron, and
Perkin Elmer. Any other sources ? We are looking for an intrument that can
do microspectroscopy, i.e., spatially resolved IR spectroscopy on ~ 1 micron
(or better, of course) length scale.

Thanks in advance
Andreas

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Andreas Taubert
Dept. of Materials Science & Engineering
3231 Walnut St.
University of Pennsylvania
Philadelphia PA 19104-6272
Tel.: 215 573 1179
Fax: 215 573 2128
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~



From daemon Mon Jun 24 10:32:57 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Mon, 24 Jun 2002 11:21:02 -0400
Subject: TEM POSITION

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Transmission Electron Microscopist

Princeton University's Department of Molecular Biology is seeking a
Transmission Electron Microscopist to run a new LEO912AB. Duties will
include: (1) teaching and training undergraduate and graduate students,
post-docs and faculty on instrument operation, (2) preparation and
sectioning of diverse biological specimens including yeast, virally infected
mammalian cells, brain slices and Drosophila embryos, using a variety of
methods including high-pressure freezing and freeze substitution techniques,
(3) daily microscope system checks and scheduling of users, (4) general lab
maintenance and supply. The candidate should have a BS/MS in a biological
field and several years of TEM experience. Knowledge of ultra-structural
and immuno-labeling protocols, the operation and care of microtomes and
other equipment needed for specimen preparation, a basic understanding of
digital imaging and the ability to manage and administer digital
workstations and image archives is essential. The candidate must be highly
interactive, willing to collaborate on diverse projects and able to identify
and research the best methods of specimen preparation and examination. Rank
and salary are dependent upon qualifications and experience. Please send
curriculum vitae, a list of references and representative samples of your
work to Professor Mark Rose, Chair, Search Committee, Department of
Molecular Biology, Princeton University, Princeton, NJ 08544-1014.
Princeton University is an Equal Opportunity/Affirmative Action Employer



Joe Goodhouse
Confocal / EM Core Lab Manager
Department of Molecular Biology
Princeton University
jgoodhouse-at-molbio.princeton.edu
609-258-5432

Visit us at http://www.molbio.princeton.edu/facility/confocal/index.html



From daemon Mon Jun 24 10:54:34 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Mon, 24 Jun 2002 11:44:42 -0400
Subject: ignore-2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Thank you too.


Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University
609-258-5432



From daemon Mon Jun 24 11:18:01 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Mon, 24 Jun 2002 12:08:03 -0400
Subject: ignore test

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thank you.

Joe Goodhouse
Confocal / EM Core Lab Manager
Department of Molecular Biology
Princeton University
jgoodhouse-at-molbio.princeton.edu
609-258-5432

Visit us at http://www.molbio.princeton.edu/facility/confocal/index.html



From daemon Mon Jun 24 11:22:25 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Mon, 24 Jun 2002 12:13:09 -0400
Subject: ignore

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thank you

Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University



From daemon Mon Jun 24 15:51:57 2002



From: Jolanta Jacobs :      jolanta-at-ksu.edu
Date: Mon, 24 Jun 2002 16:01:21 -0500
Subject: LM chloral hydrate in Alexander's stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi everybody,
I need to do a vital stain on pollen of Arabidopsis mutant I'm working on-it
seems Alexander's stain is still widely used, at least outside US. One of
the ingredients is chloral hydrate which happens to be a schedule IV
substance. I called DEA to get the license but it may take longer than I
have the time for. Is there a substance that may be used instead of chloral
hydrate? Or some other as easy as Alexander's vital staining method for
pollen?
Thank you for any input you may have,
Jolanta Jacobs
Grad student in Div of Biology
Manhattan, Kansas



From daemon Mon Jun 24 15:54:59 2002



From: Chun-Ming Li :      lichun-at-uiuc.edu
Date: Mon, 24 Jun 2002 15:48:14 -0500
Subject: Re: TEM sample preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear friends,

Thank you very much for your useful suggestions on polishing multilayers
on glass! I will try them.

Best regards,
Chunming Li

Sousan Abolhassani wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Chun-Ming,
}
} As your multi layer is thick, in any case you should thin it
} afterwards.
} Instead of going through the problem of dissolving and collecting
} such a layer that you still need to thin by other techniques
} there are at least three possibilities (that do not need suffisticated
} facilities):
}
} a. If you want to see the top view of your multi-layer, then
} it means that you should thin your sample from the molybdenum
} side until it is electron transparent. For this, the sample
} should be grinded from the glass side to a thickness of the order
} of 20-50 microns, then mounted on a ring sample holder used
} for ion milling. (If you can grind down to 30 micrometers it needs a
} shorter time for ion-milling, but if you cann't it is not important).
} Then you ion-mill your sample from the glass side, until you see a hole
} and you know that the thinnest part is in the CuInSe2 region.
}
} Note that in this case you will not see the molybdenum, because your
} CuInSe2 is 0.7-2.5 micrometers as you say.
} If you wish to observe the molybdenum as well, you could take another
} sample and proceed as above, but stop milling when the sample is
} near the Mo (you could see from colour changes) and then ion mill
} from the other side.
}
} b. The second possibility is to do a tripod polishing and this can
} be done with different orientations. All layers will be observable
} and from my experience this is the most convenient method, that
} does not need to complicated facilities.
}
} c. The last method, that I use for quick determinations of the
} material before the above elaborate methods, is to scratch the
} surface to be observed by a diamond pointer and then put the
} sctratched debris in a solvent that does not attack the film
} (you should use a very small recipient not to loose your material)
} and collect the debris on a grid with a thin support film.
} This method is very rapid and can quickly bring some information
} before you proceed with more elaborate techniques.
}
} Hope this helps,
}
} Sousan
}
} E-mail: sousan.abolhassani-at-psi.ch
} http://www.psi.ch
}
} Chun-Ming Li wrote:
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} } Dear friends:
} } I have a problem in preparing a TEM sample (plan view). The
} } as-received sample is as follow:
} } On a soda-lime glass (thickness 2-4 mm) substrate, deposit a
} } molybdenum layer (thickness 0.1-1.5 micron), and then deposit a layer of
} } CuInSe2 (thickness 0.7-2.5 micron).
} } After dissolved the glass with HF (hydrofluoric) acid (cover the layers
} } with lacquer before etching, otherwise the multilayers will also be
} } dissolved), I immersed the multilayers into acetone. But the layers are
} } too brittle that they broke into tiny chips. Even sometimes I can get a
} } big chip, it is still very difficult to fix it on a grid, since it is
} } too brittle.
} } I would appreciate it very much if you can give me any suggestions
} } to solve this problem, i.e., how to get a plan-view TEM sample from the
} } as-received sample. The major problem would be how to get rid of the
} } glass and fix the multilayers to grid before ion milling.
} } Thank you in advance!
} }
} } Chun-Ming Li, Ph.D.
} } Dept. of Mater. Sci. and Engi.
} } University of Illinois at Urbana-Champaign



From daemon Mon Jun 24 17:07:50 2002



From: Jolanta Jacobs :      jolanta-at-ksu.edu
Date: Mon, 24 Jun 2002 17:18:22 -0500
Subject: LM chloral hydrate in Alexander's stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi everybody,
I need to do a vital stain on pollen of Arabidopsis mutant I'm working on-it
seems Alexander's stain is still widely used, at least outside US. One of
the ingredients is chloral hydrate which happens to be a schedule IV
substance. I called DEA to get the license but it may take longer than I
have the time for. Is there a substance that may be used instead of chloral
hydrate? Or some other as easy as Alexander's vital staining method for
pollen?
Thank you for any input you may have,
Jolanta Jacobs
Grad student in Div of Biology
Manhattan, Kansas



From daemon Mon Jun 24 20:30:03 2002



From: Paul Fischione :      pe_fischione-at-fischione.com (by way of
Date: Mon, 24 Jun 2002 20:20:38 -0500
Subject: Microscopy Society of America (MSA) Needs You

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} From: Microscopy Society of America: Marketing and Communications Committee
(MSAMCC)

Colleagues:

MSA is a vibrant society which encompasses all facets of Microscopy
and Microanalysis. The society and it's members are dedicated to the
promotion and advancement of the knowledge of the science and practice
of all microscopical imaging, analysis and diffraction techniques useful
for elucidating the ultrastructure and function of materials in diverse
areas of biological, materials, medical and physical sciences. To this end
MSA is a resource which is second to none throughout the world. If you're
not currently a member of MSA, then you're missing out on both the
opportunities and the fun; and we would like to take this opportunity to
invite you to join our family.


If you are already an MSA member, take a moment to talk to your colleagues
and/or students about MSA's great benefits and encourage them to join the
society. Invite them to be come part of our world wide microscopy community.


MSA Membership Benefits Include:

* A subscription to the high-quality journal: Microscopy & Microanalysis
-included with your membership dues
http://www.msa.microscopy.com/MM/MscopyManalysis.html
* A complimentary copy of Microscopy Today
http://microscopy-today.com
* Receive advanced information and reduced fees for the Annual Meeting
- the preeminent microscopy related meeting and exhibition
http://www.microscopy.com/MSAMeetings/MMMeeting.html
* Access to extensive educational materials & activities
http://www.msa.microscopy.com/RefEdu.html
* On-line Directory of MSA members
http://www.msa.microscopy.com/MSAMembers/SearchMembersDB.html
* Opportunities to have your accomplishments recognized as an Awards recipient.
http://www.msa.microscopy.com/MSADocs/MSAAwards.html
* MSA Placement Office
http://www.msa.microscopy.com/PlacementOffice/JobListings.html
* MSA Certification Program
http://www.cvmbs.colostate.edu/emcenter/msa/certboard/
* MSA Technologist Forum
http://www.cvmbs.colostate.edu/emcenter/msa/techforum/
* MSA Public Policy Committee
-remain aware and influence governmental funding opportunities
http://www.msa.microscopy.com/PublicPolicy/
* MSA Sustaining Members Program
http://www.msa.microscopy.com/cgi-bin/SMDBListing.pl
* MSA Vendor News and Surplus Equipment Listings
http://www.msa.microscopy.com/News/NewsListings.html
http://www.msa.microscopy.com/SurplusEquipment/SurplusListings.html
* Reduced prices for books and journals
* Subscription to the MSA Bulletin to keep abreast of Society activities
* Specialized networking via participation in "Focussed Interest Groups" with
microscopist's having similar interests to yours
* and of course the International Networking facilitated in part by
the Microscopy
Listserver to keep abreast of worldwide microscopy activities
and information.

So if your not a member of MSA, take a moment and join!
It is easy and inexpensive [$15/Students, $ 45/Regular,
$63/outside of North America and don't forget this includes the
a subscription to the Journal].

You can become a Member On-Line at
http://www.msa.microscopy.com/MSADocs/MSANewForm.html
or request that an printed application form be sent to you by
sending an Email to:
MSABusinessOffice-at-msa.microscopy.com


Remember MSA needs YOU !
--------------------

MSA Marketing and Communications Committee
MSAMCC-at-msa.microscopy.com


From daemon Mon Jun 24 20:30:09 2002



From: Brendan J. Griffin :      bjg-at-cyllene.uwa.edu.au (by way of
Date: Mon, 24 Jun 2002 20:18:36 -0500
Subject: CMM Biological Microscopy Lectureship - open til 5th July

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear All,

Please note that the closing date for the advertised Biological
Microscopy Lectureship is 5th July, not 28th June.

Thank you
--




Brendan J. Griffin
Director & Assoc. Professor in Electron Microscopy
Centre for Microscopy and Microanalysis
The University of Western Australia
First floor, Physics Building,
35 Stirling Highway
CRAWLEY, WA, AUSTRALIA 6009
ph 61-8-9380-2739 fax 9380-1087
mobile 0409-104-096
bjg-at-cmm.uwa.edu.au
http://cmm.uwa.edu.au/

CRICOS Provider No. 00126G


From daemon Tue Jun 25 10:36:34 2002



From: Deborah Grimm :      dgrimm-at-tulane.edu (by way of MicroscopyListserver)
Date: Tue, 25 Jun 2002 10:16:52 -0500
Subject: TEM Position - Tulane University

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Position Laboratory Supervisor

Tulane University's Coordinated Instrumentation Facility seeks an individual
to supervise its expanding microscopy facility. The successful candidate
will have a B.S. in a basic science and two years of electron or advanced
microscopy experience. Experience working in a multi-user environment
serving biological and materials science clientele is a plus. Direct
experience with JEOL's SEMs and TEMs as well as confocal microscopy is
desired. The position is responsible for sample preparation, instrument
maintenance and operation.

Deborah A. Grimm, Ph.D.
Director
Coordinated Instrumentation Facility
605 Lindy Boggs Building
Tulane University
New Orleans, LA 70118


From daemon Tue Jun 25 12:07:41 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Tue, 25 Jun 2002 12:40:44 -0400
Subject: Bill Miller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm trying to contact Bill Miller. Could someone send me Bill Millers email
or contact info.

Bill, you sent email to me yesterday about new high res image capture
system. I've lost it. Please contact me again.

Thanks and Regards,


Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University
609-258-5432

Visit us at http://www.molbio.princeton.edu/facility/confocal/index.html



From daemon Tue Jun 25 16:44:17 2002



From: John C. Wheatley :      John.Wheatley-at-asu.edu
Date: Tue, 25 Jun 2002 14:35:52 -0700
Subject: Position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Post-Doctoral Associate

Available with the Center for High Resolution Electron Microscopy at
Arizona State University. Position is sponsored by major chemical company
and primary focus of the research will be the development and application
of environmental transmission electron microscopy to industrially relevant
catalyst systems. Areas of particular interest include the study of phase
transformations under reaction environments, in situ sintering, mobility
and dynamic microstructural changes. Candidate will have a Ph.D. in
material science, material physics, solid-state chemistry or chemical
engineering, with experience in nano-particle characterization by in situ
transmission electron microscopy. Experience with the techniques of high
resolution imaging, electron energy loss spectroscopy and scanning
transmission electron microscopy is preferred. Please submit your resume
together and the names and addresses of 3 referees to: Dr. Peter A.
Crozier, Industrial Associates Program, Center for Solid State Science,
Arizona State University, Tempe, AZ 85287-1704, Fax (480) 965-9004, email
crozier-at-asu.edu.

John C. Wheatley
Lab Manager
Arizona State University
Center for Solid State Science
PSA-213
BOX 871704
Tempe, AZ 85287-1704


Phone: (480) 965-3831
FAX: (480) 965-9004
John.Wheatley-at-ASU.Edu




From daemon Tue Jun 25 17:24:26 2002



From: Al Coritz :      Cactusgrower-at-earthlink.net
Date: Tue, 25 Jun 2002 18:17:31 -0400
Subject: Bill Miller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bill Miller can be reached at microbill-at-mohawk.net

Best,

Al Coritz, Sales Manager
Diatome USA
V: 215-646-1478
F: 215-646-8931
Cactusgrower-at-earthlink.net
www.emsdiasum.com


-----Original Message-----
} From: Goodhouse, Joseph [mailto:jgoodhouse-at-molbio.princeton.edu]
Sent: Tuesday, June 25, 2002 12:41 PM
To: Microscopy (E-mail)


I'm trying to contact Bill Miller. Could someone send me Bill Millers
email
or contact info.

Bill, you sent email to me yesterday about new high res image capture
system. I've lost it. Please contact me again.

Thanks and Regards,


Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University
609-258-5432

Visit us at
http://www.molbio.princeton.edu/facility/confocal/index.html




From daemon Wed Jun 26 04:16:02 2002



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Wed, 26 Jun 2002 11:03:52 +0200
Subject: Revisiting LaB6

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a Technai 12 which we want to convert to LaB6. Since different
microscopes have their own petty preferences (not including the
manufactures). Therefore I would like to know what other manufacturers LaB6
run nicely on a Technai 12? I know Den co was used on the CM range.

It will be a pretty expensive exercise to buy a LaB6 and find it not willing
to marry my Technai!

Mr S. H. Coetzee
Electron Microscope Unit
University of Botswana
Private Bag 0022
Gabarone
Botswana
{ {...OLE_Obj...} }

Phone : +267 355 2426
Mobile: +267 718 96 729
Fax : +267 585 097
e-mail : coetzees-at-mopipi.ub.bw {mailto:coetzees-at-mopipi.ub.bw}




From daemon Wed Jun 26 06:43:14 2002



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Wed, 26 Jun 2002 13:31:43 +0200
Subject: LaB6 revisited

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all

We have a Technai 12 which we want to upgrade with a LaB6. What we would
like to know which other brands of LaB6 marry nicely with the Technai 12. I
am interested in stability and actual life.
Thanks

Mr S. H. Coetzee
Electron Microscope Unit
University of Botswana
Private Bag 0022
Gabarone
Botswana
{ {...OLE_Obj...} }

Phone : +267 355 2426
Mobile: +267 718 96 729
Fax : +267 585 097
e-mail : coetzees-at-mopipi.ub.bw {mailto:coetzees-at-mopipi.ub.bw}




From daemon Wed Jun 26 07:57:49 2002



From: John Hoffpauir :      John.Hoffpauir-at-mail.tju.edu
Date: Wed, 26 Jun 2002 08:51:09 -0400
Subject: Position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Experienced (20+years) electron microscopist looking for a weekend opening in
a clinical/research lab in the Philadelphia area. f anyone knows of any
opening please feel free to email me.
John Hoffpauir


From daemon Wed Jun 26 08:46:33 2002



From: Lena Falk :      lklfalk-at-fy.chalmers.se
Date: Wed, 26 Jun 2002 15:42:16 +0200
Subject: SEM course at Chalmers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear fellow microscopist,

Take the chance to learn about the developments in scanning electron
microscopy! A course on "Scanning Electron Microscopy - Imaging and
Microanalysis", SEM 2002, will be given at Chalmers University of
Technology, Göteborg, Sweden, October 22 - 24, 2002.

The course will discuss imaging and microanalysis in high vacuum, low
vacuum and environmental SEM’s with special focus on:
o low voltage SEM
o new detectors and detector technology
o electron sources
o elemental and phase analysis
o optimum specimen preparation

The course will be given in close collaboration with microscope
manufacturers and suppliers of equipment for microanalysis.
Instrumentation from the following companies will be used in the course:
SEM’s from Fei, Hitachi, Jeol and Leo, EDX/WDX/EBSP systems from EDAX,
Oxford Instruments and HKL Technology, and equipment for specimen
preparation from Gatan.

This three-day intensive course will give a theoretical background in
the morning sessions and experimental insights in the laboratory
afternoon sessions. The lectures and the demonstrations will be given
by application specialists from the different companies representad at
the course, and also by researchers working with different applications
in SEM. The demonstrations and lab classes will be carried out on
equipment brought to Chalmers specifically for this course.

Detailed information about SEM 2002, including course programme and
registration form, is posted on: http://fy.chalmers.se/microscopy This
web-site will be continuously updated until the course starts. The
registration fee is 6.500 SEK plus VAT, including lecture notes,
refreshments, lunches and course dinner.

Looking forward to seeing you at SEM 2002!

Lena Falk and Mats Halvarsson
Course Organisers

____________________________

Associate Professor Lena Falk
Department of Experimental Physics,
Chalmers University of Technology,
SE-412 96 Göteborg, SWEDEN

tel: +46 31 772 3321
fax: +46 31 772 3224
e-mail: lklfalk-at-fy.chalmers.se


From daemon Wed Jun 26 11:03:34 2002



From: Ekstrom, Harry :      harry.ekstrom-at-honeywell.com
Date: Wed, 26 Jun 2002 12:41:10 -0700
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Please note that the entry for the Expert's Session: Core facility managment is incorrect in the program. Although it is listed for 3:00 Wed afternoon, the correct time is 8:00 Wed morning.
The program should have read as follows:

Maintaining Major Equipment in a Core Microscopy Facility

8:15 Introduction

8:30: Servicing by Original Equipment manufacturers: the structure, function & considerations for assembling & operating a major service organization
JEOL: Patrick McGinley
National Service Manager
FEI: Mike Kearney
Director of Service
Hitachi: Greg Rigby
Director of Service

9:15: Servicing EMs by On-site staff: Insight into the Do-it-yourself Approach.
John Wheatley, Arizona State University
Owen Mills, Michigan Technological University

10:00: Break

10:30: Using 3rd Party Service Organizations for Major Equipment Maintenance
Art McCanna
Service Manager
Materials Analytical Services

11:00: Policies Regarding Allowable Costs for Equipment Maintenance on National Science Foundation (NSF) Awards


--------------------------------------





Harry Ekstrom
Materials Laboratory

*Phone: (602) 231-2744
?Fax: (602) 231-1547
*e-mail: harry.ekstrom-at-honeywell.com
{mailto:harry.ekstrom-at-honeywell.com}




From daemon Thu Jun 27 08:08:57 2002



From: David_Bell-at-Millipore.com (by way of MicroscopyListserver)
Date: Thu, 27 Jun 2002 07:54:28 -0500
Subject: Microtomy of Polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello All,

I am looking for a service lab in the north east that has the capability
of embedding and sectioning various types of polymer membranes. Please
contact me off list if you can provide such a service.

Thanks in advance,



David Bell
Scientist
Electron Microscopy Lab
Millipore Corporation
80 Ashby Road
Bedford, MA 01730
(781) 533-2108


From daemon Thu Jun 27 08:09:02 2002



From: reinhard rachel :      reinhard.rachel-at-biologie.uni-regensburg.de (by
Date: Thu, 27 Jun 2002 07:54:03 -0500
Subject: Re: Revisiting LaB6

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are operating a CM12 since about 11 years now,
exclusively with LaB6 filaments (in a Microbiology
environment, no EDX, no STEM, but with Cryo).
I strongly recommend to use this filament type.
If used properly, one filament can last for as long as 3 years;
in our CM12, it is only the 4th filament now.
EM is in use daily, approx. 6 - 10 hrs. Mo - Fr.
3 years of NOT changing a filament, not opening the column:
excellent vacuum! no down times -- if you sum it up,
it even might be cheaper than W-filaments.

I started with Denca; now I'm using the Kimball Physics
filament; a bit more expensive, but recommendation from
the service engineer. Runs very smoothly, no problems.
very satisfied.

if you need more info, contact me directly.

sincerely, yours,

Reinhard Rachel


From daemon Thu Jun 27 12:03:38 2002



From: ATC SEM Laboratory :      atcsem-at-earthlink.net
Date: Thu, 27 Jun 2002 12:50:33 -0400
Subject: LaB6 life expectancy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

In one of the massages I've read that LaB6 filament lasts for ~3 years. We
run our SEM 16-20 hr per day(heat 5.0-5.3) and turn down heat to 4 for the
night. We only turn off SEM for weekends. My question is how long does LaB6
filament lasts?

Thanks,
Pavel




From daemon Thu Jun 27 12:08:56 2002



From: ATC SEM Laboratory :      atcsem-at-earthlink.net
Date: Thu, 27 Jun 2002 13:04:15 -0400
Subject: LaB6 life expectancy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

In one of the massages I've read that LaB6 filament lasts for ~3 years. We
run our SEM 16-20 hr per day(heat 5.0-5.3) and turn down heat to 4 for the
night. We only turn off SEM for weekends. My question is how long does LaB6
filament lasts?

Thanks,
Pavel





From daemon Thu Jun 27 15:55:58 2002



From: ATC SEM Laboratory :      atcsem-at-earthlink.net
Date: Thu, 27 Jun 2002 16:48:23 -0400
Subject: Re: LaB6 life expectancy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks for the info.
We keep vacuum system ON at all the time, just the electron optics console
we turn OFF before weekends.
This filament is already running for almost 11 months, and I wondered how
much longer would it last. Previously, before I start working here, LaB6 was
changed in ~ 6 months period.
Thanks again for the info.

Pavel



From daemon Thu Jun 27 16:08:41 2002



From: kwalters :      kwalters-at-blue.weeg.uiowa.edu
Date: Thu, 27 Jun 2002 16:02:21 -0500
Subject: Lynx el Tissue Processor for EM samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Has anyone had experience with this automated processor of EM samples? Thanks
for any information you can provide me.

Kathy

Kathy Walters //
Central Microscopy Research Facility / /
85 EMRB / /\
University of Iowa / /\ \
Iowa City, Iowa 52242 / / \ \
Phone #: (319) 335-8142 / / \ \
Fax #: (319) 384-4469 ______ ((0))
email: Katherine-Walters-at-uiowa.edu |__| / /
|| / /
--------------
------------------



From daemon Thu Jun 27 16:58:45 2002



From: Judy Trogadis :      TrogadisJ-at-smh.toronto.on.ca
Date: Thu, 27 Jun 2002 17:06:55 -0400
Subject: facility management

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This topic may have been discussed, if so, I will look at the archives if you tell me an approximate date.

Some of the microscopes in our facility are sometimes left in a sorry state by inexperienced or hasty users. I would like to keep track of users who are actually capturing images, therefore using the computer (if fingerprint sensors exist, let me know, I can just dust the microscope knobs).

Has anyone used software that keeps a record of login ID's and hours of use? We are on NT operating systems.

Thank you

Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 8Queen
30 Bond St.
Toronto, ON
M5B 1W8
Canada

ph: 416-864-6060 x6337
fax: 416-864-6043
trogadisj-at-smh.toronto.on.ca




From daemon Fri Jun 28 06:33:35 2002



From: ATC SEM Laboratory :      atcsem-at-earthlink.net
Date: Fri, 28 Jun 2002 07:25:06 -0400
Subject: Re: LaB6 life expectancy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reading you question ,Ritchie, I thought you might be right. The
electronics probably would last longer if I don't turn on and off the
console 50 times a year, but I think TV monitors, would burn out.

Regards,
Pavel



From daemon Fri Jun 28 08:15:07 2002



From: danthony38-at-cogeco.ca (by way of Ask-A-Microscopist)
Date: Fri, 28 Jun 2002 08:04:38 -0500
Subject: Ask-A-Microscopist: Polarizer and Crystals in LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (danthony38-at-cogeco.ca) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Wednesday,
June 26, 2002 at 13:32:34
---------------------------------------------------------------------------

Email: danthony38-at-cogeco.ca
Name: don anthony

Education: 9-12th Grade High School

Location: peterborough ont canada

Question: I have seen beautiful photos of, (as I was told)chemicals
crystalized on glass slides. As it was explained to me, a polorizer
was placed in front of the light source,and a second polorizer, above
the crystals being photographed. The profusion of colour was
unbelieveable, and when the polorizer was moved slightly, the colours
changed, creating a different picture. My question is,after trying to
duplicate this proceedure, I'm getting no where, can anyone help?
Also what chemicals make for the best crystals. Many thanks Don

---------------------------------------------------------------------------


From daemon Fri Jun 28 08:15:12 2002



From: jolanta-at-ksu.edu ()
Date: Fri, 28 Jun 2002 08:03:47 -0500
Subject: Ask-A-Microscopist: stain on pollen- chloral hydrate?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (jolanta-at-ksu.edu) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html#form on Monday,
June 24, 2002 at 17:37:44
---------------------------------------------------------------------------

Email: jolanta-at-ksu.edu
Name: Jolanta Jacobs

Organization: KSU

Education: Graduate College

Location: Manhattan, KS USA

Question: Hello,
I need to do a vital stain on pollen of Arabidopsis mutant I'm working on- it
seems Alexander's stain is still widely used, at least outside US. One of
the ingredients is chloral hydrate which happens to be a schedule IV
substance. I called DEA to get the license but it may take longer than I
have the time for. Is there a substance that may be used instead of chloral
hydrate? Or some other as easy as Alexander's vital staining method for
pollen?
Thank you for any input you may have,
Jolanta Jacobs
Grad student in Div of Biology
Manhattan, Kansas


---------------------------------------------------------------------------


From daemon Fri Jun 28 08:26:26 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Fri, 28 Jun 2002 09:19:21 -0400
Subject: RE: facility management

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Morning Judy,
You need to speak to your network manager. S/he can set up an audit
of resources that will keep track of the information. Also, you should
discuss means by which users may be isolated to specific resources on the
network or local client such that their use of resources can be structured.
User Manager for Domains permits domain-wide auditing of user activities
along with auditing of specific resource use. You will NOT be able to keep
a microscope clean when the user doesn't pay, however. If entry to the
facility/office is by pass card, then there will be a record of entry and
times. Perhaps you might consider placing a sign on the scope that
requests that any user report any scope in poor condition so that you know,
at least, that s/he wasn't the culprit. If you know the previous user, you
can gently chide. For a student, the grade's the thing.
Such things are easier the more widespread is use of NT Wks clients
in your system. If Win 9x is the majority client OS, then you will have to
rely more on domain auditing at the server. NT Wks permits local auditing.
What is called "shell scripting" is the method of choice to specify and
amplify auditing requirements, especially for specific resources (e.g..,
shares, applications, and other objects such as printers). Such 'scripts'
are little more than advanced batch files and may be mounted on NT Wrks
clients to reduce network overhead. Your domain admins should be able to
tailor auditing to your needs. If YOU are the Admin, then you should get
some help from the System Admin.

There is a book by Minasi, Mark, "Mastering Windows NT Server 4",
Sybex, which should be found in the libraries of at least one of your IT
folks. You might also ask about another that I use. Hill, T, "Windows NT
Scripting", Macmillan.

Regards,

Fred Monson

} ----------
} From: Judy Trogadis
} Sent: Thursday, June 27, 2002 5:06 PM
} To: Microscopy-at-sparc5.microscopy.com
} Cc: Judy Trogadis
} Subject: facility management
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} This topic may have been discussed, if so, I will look at the archives if
} you tell me an approximate date.
}
} Some of the microscopes in our facility are sometimes left in a sorry
} state by inexperienced or hasty users. I would like to keep track of users
} who are actually capturing images, therefore using the computer (if
} fingerprint sensors exist, let me know, I can just dust the microscope
} knobs).
}
} Has anyone used software that keeps a record of login ID's and hours of
} use? We are on NT operating systems.
}
} Thank you
}
} Judy Trogadis
} Bio-Imaging Coordinator
} St. Michael's Hospital, 8Queen
} 30 Bond St.
} Toronto, ON
} M5B 1W8
} Canada
}
} ph: 416-864-6060 x6337
} fax: 416-864-6043
} trogadisj-at-smh.toronto.on.ca
}
}
}
}


From daemon Fri Jun 28 08:55:05 2002



From: Corazon D. Bucana :      bucana-at-audumla.mdacc.tmc.edu
Date: Fri, 28 Jun 2002 08:36:42 -0400
Subject: travel to Quebec City

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would appreciate clarification on how to get to Quebec city from
Montreal-Dorval airport. The website is not very clear on how to get to
Quebec City i.e. the shuttle service lists the airport to downtown Montreal
but I was not able to locate the shuttle to Quebec City. Or does this mean
to take the shuttle to downtown Montreal and take another shuttle to Quebec
city? Please help.

Thanks,

Cora Bucana
Corazon D. Bucana, Ph.D.
Dept. Cancer Biology
UT M.D.Anderson Cancer Center
1515 Holcombe Blvd, Box 173
Houston, Texas 77030
Phone: 713-792-8106
FAX 713-792-8747


From daemon Fri Jun 28 10:41:09 2002



From: Monson, Frederick C. :      fmonson-at-wcupa.edu
Date: Fri, 28 Jun 2002 11:32:15 -0400
Subject: RE: facility management

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Morning Judy,
You need to speak to your network manager. S/he can set up an audit
of resources that will keep track of the information. Also, you should
discuss means by which users may be isolated to specific resources on the
network or local client such that their use of resources can be structured.
User Manager for Domains permits domain-wide auditing of user activities
along with auditing of specific resource use. You will NOT be able to keep
a microscope clean when the user doesn't pay, however. If entry to the
facility/office is by pass card, then there will be a record of entry and
times. Perhaps you might consider placing a sign on the scope that
requests that any user report any scope in poor condition so that you know,
at least, that s/he wasn't the culprit. If you know the previous user, you
can gently chide. For a student, the grade's the thing.
Such things are easier the more widespread is use of NT Wks clients
in your system. If Win 9x is the majority client OS, then you will have to
rely more on domain auditing at the server. NT Wks permits local auditing.
What is called "shell scripting" is the method of choice to specify and
amplify auditing requirements, especially for specific resources (e.g..,
shares, applications, and other objects such as printers). Such 'scripts'
are little more than advanced batch files and may be mounted on NT Wrks
clients to reduce network overhead. Your domain admins should be able to
tailor auditing to your needs. If YOU are the Admin, then you should get
some help from the System Admin.

There is a book by Minasi, Mark, "Mastering Windows NT Server 4",
Sybex, which should be found in the libraries of at least one of your IT
folks. You might also ask about another that I use. Hill, T, "Windows NT
Scripting", Macmillan.

Regards,

Fred

Frederick C. Monson, PhD
Center for Advanced Scientific Imaging
Schmucker II Science Center
West Chester University
South Church Street and Rosedale
West Chester, Pennsylvania, USA, 19383
Phone: 610-738-0437
FAX: 610-738-0437
fmonson-at-wcupa.edu
CASI URL: http://darwin.wcupa.edu/casi/
WCUPA URL: http://www.wcupa.edu/
Visitors URL: http://www.wcupa.edu/_visitors/



} ----------
} From: Judy Trogadis
} Sent: Thursday, June 27, 2002 5:06 PM
} To: Microscopy-at-sparc5.microscopy.com
} Cc: Judy Trogadis
} Subject: facility management
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} This topic may have been discussed, if so, I will look at the archives if
} you tell me an approximate date.
}
} Some of the microscopes in our facility are sometimes left in a sorry
} state by inexperienced or hasty users. I would like to keep track of users
} who are actually capturing images, therefore using the computer (if
} fingerprint sensors exist, let me know, I can just dust the microscope
} knobs).
}
} Has anyone used software that keeps a record of login ID's and hours of
} use? We are on NT operating systems.
}
} Thank you
}
} Judy Trogadis
} Bio-Imaging Coordinator
} St. Michael's Hospital, 8Queen
} 30 Bond St.
} Toronto, ON
} M5B 1W8
} Canada
}
} ph: 416-864-6060 x6337
} fax: 416-864-6043
} trogadisj-at-smh.toronto.on.ca
}
}
}
}


From daemon Fri Jun 28 10:58:53 2002



From: Kevin Frischmann :      kfrisch-at-amnh.org
Date: Fri, 28 Jun 2002 11:52:07 -0400
Subject: Re: facility management (NT user auditing)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Judy,

I believe you can obtain the information you need without installing any additional software. Windows NT has built-in tools for tracking users' logons/logoffs (among other things), though it's not enabled by default.

Click Start -} Programs -} Administrative Tools (Common) -} User Manager
Select the Policies menu -} Audit..., and you'll find a window that allows you to enable auditing, and select which events are audited and logged. In your case, I'd recommend just enabling auditing for successful logons and logoffs. Everything selected will be logged to the Security Event Log: C:\WINNT\system32\config\SecEvent.Evt (Assuming %SystemRoot% is C:\WINNT\).

In order to view, filter, sort, and archive the log, as well as change settings like maximum file size and overwrite policy, use the Event Viewer:
Click Start -} Programs -} Administrative Tools (Common) -} Event Viewer

Use Log -} Security to select the Security Log for viewing. I think you'll find the View -} Filter Events... selection will allow you to do everything you need, without having to import to a spreadsheet.

Best Regards,

Kevin Frischmann

Microscopy & Imaging Facility
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 USA

Phone: 212-313-7975
Fax: 212-496-3480
email: kfrisch-at-amnh.org


At 05:06 PM 6/27/02 -0400, Judy Trogadis wrote:
} This topic may have been discussed, if so, I will look at the archives if you tell me an approximate date.
}
} Some of the microscopes in our facility are sometimes left in a sorry state by inexperienced or hasty users. I would like to keep track of users who are actually capturing images, therefore using the computer (if fingerprint sensors exist, let me know, I can just dust the microscope knobs).
}
} Has anyone used software that keeps a record of login ID's and hours of use? We are on NT operating systems.
}
} Thank you
}
} Judy Trogadis
} Bio-Imaging Coordinator
} St. Michael's Hospital, 8Queen
} 30 Bond St.
} Toronto, ON
} M5B 1W8
} Canada
}
} ph: 416-864-6060 x6337
} fax: 416-864-6043
} trogadisj-at-smh.toronto.on.ca



From daemon Fri Jun 28 11:12:21 2002



From: Kevin Frischmann :      kfrisch-at-amnh.org
Date: Fri, 28 Jun 2002 12:05:28 -0400
Subject: Re: facility management (NT user auditing)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Judy,

I believe you can obtain the information you need without installing any additional software. Windows NT has built-in tools for tracking users' logons/logoffs (among other things), though it's not enabled by default.

Click Start -} Programs -} Administrative Tools (Common) -} User Manager
Select the Policies menu -} Audit..., and you'll find a window that allows you to enable auditing, and select which events are audited and logged. In your case, I'd recommend just enabling auditing for successful logons and logoffs. Everything selected will be logged to the Security Event Log: C:\WINNT\system32\config\SecEvent.Evt (Assuming %SystemRoot% is C:\WINNT\).

In order to view, filter, sort, and archive the log, as well as change settings like maximum file size and overwrite policy, use the Event Viewer:
Click Start -} Programs -} Administrative Tools (Common) -} Event Viewer

Use Log -} Security to select the Security Log for viewing. I think you'll find the View -} Filter Events... selection will allow you to do everything you need, without having to import to a spreadsheet.

Best Regards,

Kevin Frischmann

Microscopy & Imaging Facility
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 USA

Phone: 212-313-7975
Fax: 212-496-3480
email: kfrisch-at-amnh.org


} At 05:06 PM 6/27/02 -0400, Judy Trogadis wrote:
} This topic may have been discussed, if so, I will look at the archives if you tell me an approximate date.
}
} Some of the microscopes in our facility are sometimes left in a sorry state by inexperienced or hasty users. I would like to keep track of users who are actually capturing images, therefore using the computer (if fingerprint sensors exist, let me know, I can just dust the microscope knobs).
}
} Has anyone used software that keeps a record of login ID's and hours of use? We are on NT operating systems.
}
} Thank you
}
} Judy Trogadis
} Bio-Imaging Coordinator
} St. Michael's Hospital, 8Queen
} 30 Bond St.
} Toronto, ON
} M5B 1W8
} Canada
}
} ph: 416-864-6060 x6337
} fax: 416-864-6043
} trogadisj-at-smh.toronto.on.ca



From daemon Fri Jun 28 11:53:46 2002



From: Kevin Frischmann :      kfrisch-at-amnh.org
Date: Fri, 28 Jun 2002 12:45:13 -0400 (EDT)
Subject: Re: facility management

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Judy,

I believe you can obtain the information you need without installing any
additional software. Windows NT has built-in tools for tracking users'
logons/logoffs (among other things), though it's not enabled by default.

Click Start -} Programs -} Administrative Tools (Common) -} User Manager
Select the Policies menu -} Audit..., and you'll find a window that allows
you to enable auditing, and select which events are audited and logged.
In your case, I'd recommend just enabling auditing for successful logons
and logoffs. Everything selected will be logged to the Security Event
Log: C:\WINNT\system32\config\SecEvent.Evt (Assuming %SystemRoot% is
C:\WINNT\).

In order to view, filter, sort, and archive the log, as well as change
settings like maximum file size and overwrite policy, use the Event
Viewer:
Click Start -} Programs -} Administrative Tools (Common) -} Event Viewer

Use Log -} Security to select the Security Log for viewing. I think
you'll find the View -} Filter Events... selection will allow you to do
everything you need, without having to import to a spreadsheet.

This is the easiest to configure option I can think of; If you need
something a bit more sophisticated, and/or need to audit a large number of
workstations, Fred Monson's suggestions are certainly more appropriate.

Best Regards,

Kevin Frischmann

Microscopy & Imaging Facility
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 USA

Phone: 212-313-7975
Fax: 212-496-3480
email: kfrisch-at-amnh.org




At 05:06 PM 6/27/02 -0400, Judy Trogadis wrote:
This topic may have been discussed, if so, I will look at the archives if
you tell me an approximate date.

Some of the microscopes in our facility are sometimes left in a sorry
state by inexperienced or hasty users. I would like to keep track of users
who are actually capturing images, therefore using the computer (if
fingerprint sensors exist, let me know, I can just dust the microscope
knobs).

Has anyone used software that keeps a record of login ID's and hours of
use? We are on NT operating systems.

Thank you

Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 8Queen
30 Bond St.
Toronto, ON
M5B 1W8
Canada

ph: 416-864-6060 x6337
fax: 416-864-6043
trogadisj-at-smh.toronto.on.ca




From daemon Fri Jun 28 20:54:16 2002



From: rcmoretz-at-att.net
Date: Sat, 29 Jun 2002 01:41:12 +0000
Subject: Re: Lynx el Tissue Processor for EM samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Kathy:
I have had experience with the Lynx processor for almost
as long as they have been on the market--first directly,
then thru Leica. For routine processing of tissues, the
Lynx offers the opportunity to do a one time setup for
processing from fixative to epoxy:dehydrant mix. I
currently have programs for OsO4 thru pure epoxy,
including uranyl acetate en bloc staining that runs for
more than 24 hours. The only caveat I have for the pure
epoxy (epon replacement) is that I run overnight up to
pure, mix and add the pure the following morning (I am
reluctant to let the complete epoxy mix stand overnight--
there's too much polymerization going on, even at RT).
The programming is simple and will allow up to 9
separate regimens (I have only 4 plus a test routine set
up), and alteration is simple to allow for different
start time, etc. The unit has a Peltier cooled chamber
that will permit 4C up to RT or higher for the different
steps. There is an agitation feature, etc. The maximum
number of samples (separate wells) is 28, altho'
chambers are available for 9 wells. It is possible to
us the Lynx for TEM grid staining--but I have to admit
that I don't trust the system for doing lead staining!
So, I've never done that.
All is not always great. I had no trouble with the
original unit, which is why I purchased the second one
(13 years ago). But it was the proverbial lemon. It
took nearly 5 years to get it functional and it
generally operates flawlessly--altho' I still have
gremlin-like problems now and again to remind me that it
really ought to be a bright yellow color. I now
purchase supplies thru' EMS and use the unit for
processing large volumes of tissues for critical
studies. Proper attention to details, careful cleaning
and maintenance and not over-reusing the reagent vials
will aid in improved,consistent operational
performance. (Indeed, the last problem I had was
operator-induced, from using a warped, reused reagent
vial!) Like any other equipment, the Lynx is a
potentially great time and effort saver, but poor
routines (and the occasional lemon) can result in severe
headaches (the thing _always_ goes off at about 2am!).

Roger
--
Where the world is only slightly
less weird than it actually is.
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi,
}
} Has anyone had experience with this automated processor of EM samples? Thanks
} for any information you can provide me.
}
} Kathy
}
} Kathy Walters //
} Central Microscopy Research Facility / /
} 85 EMRB / /\
} University of Iowa / /\ \
} Iowa City, Iowa 52242 / / \ \
} Phone #: (319) 335-8142 / / \ \
} Fax #: (319) 384-4469 ______ ((0))
} email: Katherine-Walters-at-uiowa.edu |__| / /
} || / /
} --------------
} ------------------
}
}


From daemon Sat Jun 29 10:35:51 2002



From: Caroline Schooley :      schooley-at-mcn.org
Date: Sat, 29 Jun 2002 08:14:24 -0700
Subject: Re: Ask-A-Microscopist: Polarizer and Crystals in LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

We need to know more about what you've tried. What polarizers are you
using? When you look thru the pair & rotate one, does the field go really
dark, or just a bit more grey? What did you crystallize? Is it a thin
layer? These websites might help:
http://www.crystal-land.com , http://www.pbrc.hawaii.edu/~kunkel/,
http://www.cellsalive.com, http://home.att.net/~seberhard,
http://www.ScienceArt.nl

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460