From: John Arnott :      ladres-at-worldnet.att.net
Date: Mon, 30 Sep 2002 08:45:44 -0400
Subject: More thoughts on Moly vs. Pt apertures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Just some additional thoughts on material choice for apertures:

When you speak of a Pt aperture it is probably Pt/Ir. The iridium allows
for easier machining. We do pure platinum and gold apertures but only when
the application demands it.

Another important consideration is the number and sizes of a multi-hole
strip or plate. We have put up to 30 holes, some as small as 1micron, in a
single part. Since the manufacture of burr-free, concentric apertures
requires that the holes be made individually, more holes and smaller sizes
result in more material losses during production. Platinum costs can add
up.

For ultra-small holes and multi-hole parts moly should be used whenever
possible.

John Arnott
Ladd Research
83 Holly Court
Williston, VT 05495

On-line Catalog: http://www.laddresearch.com

tel: 1-802-658-4961(anywhere) or 1-800-451-3406(US)
fax: 1-802-660-8859
e-mail: sales-at-laddresearch.com




From daemon Mon Sep 30 11:29:36 2002



From: Heeschen, Bill (WA) :      WAHeeschen-at-dow.com
Date: Mon, 30 Sep 2002 12:11:45 -0400
Subject: Position Available - polymer electron microscopist at The Dow Che

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


} Please send resumes and direct inquiries to: rccieslinski-at-dow.com
}
} POSITION OPEN
}
} Polymer Electron Microscopist
} The Dow Chemical Company
} Midland, Michigan
}
} Dow’s Corporate R&D Analytical Science Laboratory has a professional level
} opening for a polymer microscopist at it’s Midland Michigan location. The
} primary responsibilities would be to design, conduct and interpret
} electron microscopy experiments to provide key answers to questions
} relating to the development, processing and the performance of various
} materials. Good written and oral communication skills are essential. The
} ability to work both independently and in a team environment is also
} extremely important.
}
} Educational Requirements:
} A Ph.D. or MS degree in a materials or chemical related discipline is
} preferred, but candidates with in a biological science related discipline
} and substantial materials EM work experience would be considered.
}
} Experience Requirements:
} Experience operation of a transmission electron microscope; excellent
} manual dexterity and visual acuity; demonstrated ability to acquire and
} objectively interpret data, and hands-on experience in the preparation of
} specimens for Electron Microscopy, including ultramicrotomy and
} cryo-ultramicrotomy.
}
} Key responsibilities will include:
} * Strong problem solving skills.
} * Active participation in development and project work teams.
} * Interpretation and documentation of work
} * Compliance with safety and quality programs.
}
} Send resumes and direct inquires to: rccieslinski-at-dow.com
}
} Robert C. Cieslinski, Ph.D.
} The Dow Chemical Company
} Corporate R&D, Analytical Sciences
} 1897E Bldg.
} Midland, MI 48667
} Fax: (989) 638-6443
} Email: rccieslinski-at-dow.com
}
} The Dow Chemical Company is the second largest chemical company in the
} world and a leader in science and technology solutions. Come visit Dow
} careers on the web at http://www.dow.com/careers.
}


From daemon Tue Oct 1 09:02:18 2002



From: Krzysztof Jan Huebner :      hubner-at-IOd.krakow.pl
Date: Tue, 1 Oct 2002 15:01:50 +0200 (MET DST)
Subject: SCIA 2003 - Call for Papers (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html





Call for Papers
13th Scandinavian Conference on Image Analysis
SCIA 2003
Göteborg, Sweden, June 29 - July 2, 2003
http://www.hh.se/scia2003

Please notice the new date

The ambition of SCIA is to bring together leading researchers in the
field of image analysis and present an exciting and advanced scientific
program reflecting the state-of-the-art of the discipline. SCIA 2003 is
the thirteenth in a series of conferences arranged every two years. We
look forward to your contributions to this long-standing tradition.

Hosted this year by the Swedish Society of Image Analysis (SSAB), the
conference is sponsored by the International Association for Pattern
Recognition (IAPR) and the Nordic chapters of IAPR.

Scientific Program

The scientific program will include contributed as well as invited
papers. Accepted papers will be published in the conference proceedings
(Springer LNCS). The first day of the conference will consist of special
sessions, such as workshops and/or tutorials.

Contributions covering the topics below are solicited. However, all
submissions addressing issues related to image analysis are strongly
encouraged. The main technical areas are:

- Image feature extraction
- Image understanding
- Grouping and segmentation
- Motion analysis
- Texture analysis
- Color analysis
- Shape analysis
- Computer Vision
- Cognitive vision
- Medical image processing
- Measurement and quantification
- Measurement and visualization
- Image coding and compression
- Multi-modal processing
- Indexing and databases
- Images and the 3-D geometry
- Standards and best practices
- Images and pattern recognition
- Classification
- Applications

Invited Speakers and Workshop/Tutorial Presenters (not yet complete)
Ivar Austvoll, Stavanger University College (NO)
Ewert Bengtsson, Uppsala University (SE)
Lars Bååth, Halmstad University (SE)
Herve Delingette, INRIA, Sophia-Antipolis (FR)
Chris Glasbey, Biomathematics & Statistics Scotland (UK)
Ed Hancock, University of York (UK)
Ioannis Kakadiaris, University of Houston (US)
Rasmus Larsen, Technical University of Denmark (DK)
Jussi Parkkinen, University of Joensuu (FI)
Milan Sonka, University of Iowa (US)

Venue
Situated in the heart of Scandinavia, Göteborg is within easy reach by
air, rail and sea. There are daily flights from all the main European
airports to Göteborg's Landvetter airport. The city center is only 20
minutes away by coach. Founded in 1829, Chalmers University of
Technology is named after the major benefactor, William Chalmers, one of
the directors of the successful Swedish East India Company in Göteborg.
Today, the campus area with its modern conference facilities is situated
in the city center close to hotels, restaurants, theatres, and shopping
centers.

Social Activities
Besides the conference dinner, there will be plenty of opportunities for
social activities including boat cruises in the beautiful Göteborg
archipelago, Liseberg - the largest amusement park in Scandinavia, the
Botanic Garden, and many others.

Instruction to Authors
Max 6 pages according to Springer LNCS format
http://www.springer.de/comp/lncs/authors.html
Look under the heading "Proceedings and Other Multi-author Volumes"
describing format files and providing help. Papers should be submitted
to:
scia2003.papers-at-hh.se

Important Dates
Submission of papers: January 22, 2003
Notification of acceptance: March 22, 2003
Camera-ready papers: May 8, 2003
Conference dates: June 29 - July 2, 2003

Contact Information
http://www.hh.se/scia2003
SCIA2003-at-gbg.congrex.se

Conference Co-Chairs
Josef Bigun
Tomas Gustavsson

Programme Committee
Fritz Albregtsen (NO)
Kalle Åström (SE)
Ivar Austvoll (NO)
Ewert Bengtsson (SE)
Ketil Bo (NO)
Magnus Borga (SE)
Gunilla Borgefors (SE)
Stefan Carlsson (SE)
Henrik Christensen (SE)
Per-Erik Danielsson (SE)
Bjarne Ersböll (DK)
Robert Forchheimer (SE)
Eric Granum (DK)
Anders Heyden (SE)
Jukka Iivarinen (FI)
Peter Johansen (DK)
Fredrik Kahl (SE)
Hans Knutsson (SE)
Björn Kruse (SE)
Rasmus Larsson (DK)
Reiner Lenz (SE)
Tony Lindeberg (SE)
Claus Madsen (DK)
Henning Nielsen (DK)
Ingela Nyström (SE)
Erkki Oja (FI)
Sören Olsen (DK)
Matti Pietikäinen (FI)
Ann Sohlberg(NO)
Örjan Smedby (SE)
Antanas Verikas (SE)
Qin Zhong-Ye (SE)





From daemon Wed Oct 2 01:03:22 2002



From: little-at-earthtech.org (by way of MicroscopyListserver)
Date: Wed, 2 Oct 2002 00:51:28 -0500
Subject: Ask-A-Microscopist: SEM: Cambridge Stereoscan questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (little-at-earthtech.org) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Tuesday,
October 1, 2002 at 21:27:21
---------------------------------------------------------------------------

Email: little-at-earthtech.org
Name: George Luce and Scott Little

Organization: Earthtech International, Inc.

Education: Graduate College

Location: Austin, Texas

Question: We have recently revived a Cambridge Stereoscan 150 SEM
from about 1978. The machine is now operational, but we cannot get
good resolution or focus above a few thousand X magnification.

We would like to communicate with someone familiar with this or a
similar model, who could give us some guidance on improving the
resolution. Details of our efforts can be supplied. We are
technically competent people with backgrounds in physics,
engineering, high voltage, electronics, vacuum, etc.

Thank you.


George Luce (geolucetx-at-aol.com)

Scott Little (little-at-earthtech.org)


Earthtech International, Inc.
Austin, Texas

Phone (512) 342-2185

www.earthtech.org

---------------------------------------------------------------------------


From daemon Wed Oct 2 01:03:27 2002



From: mgrace-at-fit.edu (by way of MicroscopyListserver)
Date: Wed, 2 Oct 2002 00:53:13 -0500
Subject: Ask-A-Microscopist:ESEM Low Vacuum Mode. how useful is it?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (mgrace-at-fit.edu) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Tuesday,
October 1, 2002 at 15:16:59
---------------------------------------------------------------------------

Email: mgrace-at-fit.edu
Name: Michael Grace

Organization: Florida Institute of Technology

Education: Graduate College

Location: Melbourne, FL, USA

Question: I am working toward aquisition of a new SEM that will see
use mostly in biological sciences applications, but will also see
some use from materials scientists, chemists, etc. I am considering
one of the new ESEMs (specifically the FEI/Philips Quanta 200.
Question: how useful is the low vacuum mode, esp. for biological
applications? Also interested in feedback from those with other
applications, but I mainly want to be sure that the money is spent
wisely- that we are not getting all the bells and whistles simply
because they exist.

Thank you.

---------------------------------------------------------------------------


From daemon Wed Oct 2 05:18:34 2002



From: Allen Sampson :      ars-at-sem.com
Date: Wed, 2 Oct 2002 05:13:21 -0700
Subject: RE: Ask-A-Microscopist: SEM: Cambridge Stereoscan questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Yes, but are you experienced in electron microscopes? This is definitely a
field that not only encompasses those mentioned, but uses them, as well as
others, all in synergistic manner.

When it comes to resolution, look to your friend, the condenser control (or
spot size or beam current or whatever else it may be called). On that
vintage of Cambridge, I expect it is called spot size. If everything else
is OK, the condenser is the primary control that determines the resolution
of the instrument.

This is because the primary aberration in an electron optic instrument is
chromatic aberration. That is the result of the wide spread in energy
ranges of the electrons emitted from the electron gun. That spread creates
what is known as a spherical aberration that affects the beam in both the X
and Y directions. As they travel through the various electromagnetic
lenses of the SEM, electrons of different energies are affected differently
primarily resulting in a different focal point in the Z axis. The
condenser lens, and associated apertures, are used to reduce this energy
spread.

Here's the deal. The higher the condenser current, the smaller the spot
size and the smaller the beam current. It's hard to explain in text, but
you actually have two condenser lenses. Their purpose is to provide a
de-magnification (reduction in apparent size) of the source of the
electrons. The source is the tungsten filament, or cathode, that emits
electrons from a rather large area of around 100 microns (micrometers).

In the process, there are a couple of imaginary images formed, or crossover
points, where apertures or electron absorbers with small central openings
are located. Electrons of different energies will be affected differently
by the electromagnetic lenses - electrons of lower energy will be displaced
more than those of higher energies. The apertures will capture those
electrons that have an energy lower or higher than that which the lenses
focus through the apertures, thus reducing the beam current by removing
those electrons with energies outside of the nominal and reducing the spot
size by ensuring that those electrons remaining have a smaller spread of
energies.

Having said that, increase the condenser current, reduce the spot size or
reduce the beam current, whichever applies.

If you're new to electron microscopy, there is one thing that really must
be stressed - cleanliness. The smallest, microscopic, piece of dust, lint,
fiber or electrically insulating contamination can wreak havoc inside an
electron optics column. They tend to build up a negative electrical charge
under influence of the beam that can create repulsive fields that deflect
the electron beam and create weird aberrations. Contamination from
fingerprints and solvents is extremely minor compared to the effects of a
single fiber in the electron path.

Lastly, what ultimately determines the resolution possible in an instrument
is the care with which it was installed. That includes the siting of the
instrument in regards to mechanical vibrations in the building as well as
the sources of electromagnetic interference. But most important is the
careful routing of the various wires and tubing within and outside the
instrument. The electron optics column is mounted on a table that is
isolated from vibrations from the frame that holds it. All connections
that run to this table must be carefully routed in order to provide similar
isolation. Of similar concern is the electrical grounding of the various
components of an SEM. Do not assume that because various large metal
components appear to have electrical continuity that they do.
Manufacturers typically provide for individual grounds to many components
for good reason.




Allen R. Sampson
Advanced Research Systems
317 North 4th. Street
St. Charles, Illinois 60174

phone (630) 513-7093 fax (630) 513-7092 http://www.sem.com


On Tuesday, October 01, 2002 10:51 PM, by way of MicroscopyListserver
[SMTP:little-at-earthtech.org] wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (little-at-earthtech.org) from
} http://www.msa.microscopy.com/Ask-A-Microscopist.html on Tuesday,
} October 1, 2002 at 21:27:21
}
---------------------------------------------------------------------------
}
} Email: little-at-earthtech.org
} Name: George Luce and Scott Little
}
} Organization: Earthtech International, Inc.
}
} Education: Graduate College
}
} Location: Austin, Texas
}
} Question: We have recently revived a Cambridge Stereoscan 150 SEM
} from about 1978. The machine is now operational, but we cannot get
} good resolution or focus above a few thousand X magnification.
}
} We would like to communicate with someone familiar with this or a
} similar model, who could give us some guidance on improving the
} resolution. Details of our efforts can be supplied. We are
} technically competent people with backgrounds in physics,
} engineering, high voltage, electronics, vacuum, etc.
}
} Thank you.
}
}
} George Luce (geolucetx-at-aol.com)
}
} Scott Little (little-at-earthtech.org)
}
}
} Earthtech International, Inc.
} Austin, Texas
}
} Phone (512) 342-2185
}
} www.earthtech.org
}
}
---------------------------------------------------------------------------
}



From daemon Wed Oct 2 05:50:11 2002



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Wed, 2 Oct 2002 11:42:04 +0100 (GMT Daylight Time)
Subject: Re: Ask-A-Microscopist:ESEM Low Vacuum Mode. how useful is it?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


In this case the "bells and whistles" are (still, I
believe) unique as they allow you to look at fully hydrated
specimens and have water in the chamber.

I am a happy XL30 ESEM (tungsten) user. If you can afford
it the FEGESEM has advantages over the tungsten filament
ESEM related to kV, spot size, SE signal and beam damage
for viewing biological samples.

Dave


On Wed, 2 Oct 2002 00:53:13 -0500 by way of
MicroscopyListserver {mgrace-at-fit.edu} wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (mgrace-at-fit.edu) from
} http://www.msa.microscopy.com/Ask-A-Microscopist.html on Tuesday,
} October 1, 2002 at 15:16:59
} ---------------------------------------------------------------------------
}
} Email: mgrace-at-fit.edu
} Name: Michael Grace
}
} Organization: Florida Institute of Technology
}
} Education: Graduate College
}
} Location: Melbourne, FL, USA
}
} Question: I am working toward aquisition of a new SEM that will see
} use mostly in biological sciences applications, but will also see
} some use from materials scientists, chemists, etc. I am considering
} one of the new ESEMs (specifically the FEI/Philips Quanta 200.
} Question: how useful is the low vacuum mode, esp. for biological
} applications? Also interested in feedback from those with other
} applications, but I mainly want to be sure that the money is spent
} wisely- that we are not getting all the bells and whistles simply
} because they exist.
}
} Thank you.
}
} ---------------------------------------------------------------------------
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



From daemon Wed Oct 2 08:01:14 2002



From: Greg Erdos :      gwe-at-biotech.ufl.edu
Date: Wed, 02 Oct 2002 08:47:06 -0400
Subject: Confocal list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Has the confocal listserver been closed or have I been purged? Haven't
gotten anything for a while.
Gregory W. Erdos, Ph.D.
Assistant Director, Biotechnology Program
Scientific Director, ICBR EM CORE
University of Florida Ph. 352-392-1295
PO Box 118525 Fax 352-846-0251
Gainesville, FL 32611 http://www.biotech.ufl.edu/EM


From daemon Wed Oct 2 09:33:14 2002



From: Anjeanette Ormonde :      Anjeanette.Ormonde-at-unilever.com
Date: Wed, 2 Oct 2002 09:13:58 -0500 (Central Daylight Time)
Subject: Confocal Microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I was hoping someone could recommend some manufacturers of confocal
microscopes.
We are just beginning to investigate the possibility of purchasing one for
our labs and while the literature I've read is intriguing, I would like
more specific information about the various scopes available and what they
can really do for us. Also, I need some idea of the price. I would
appreciate it if any manufacturers or people with personal accolades for a
specific scope would please contact me offline. Thanks for your help!

Angie



From daemon Wed Oct 2 10:55:17 2002



From: Chuck.Butterick-at-degussa.com
Date: Wed, 2 Oct 2002 10:43:21 -0500
Subject: Wood Identification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Listers,

Would someone please direct me to some resource, a good text, or better
yet, a web site that can help me identify wood types by either cross-grain
and/or with grain using a microscope? I have some crushed slivers that I'd
like to identify. Any help would be appreciated. Thanks in advance.

Chuck Butterick
Degussa Engineered Carbons
Borger, TX



From daemon Wed Oct 2 11:21:22 2002



From: Kestutis Smalinskas :      smalinskas-at-yahoo.com
Date: Wed, 2 Oct 2002 09:14:06 -0700 (PDT)
Subject: Re: Ask-A-Microscopist:ESEM Low Vacuum Mode. how useful is it?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I have a JEOL JSM-5800LV, which is a low vacuum model.
I use the LV feature quite a bit.

The down side is: you must use backscatter detection
in this mode, a slight loss in high-mag resolution,
and cost.

The plus side is convenience - samples don't have to
be coated...they can be imaged as-is. There's no
spurious EDS peaks from conductive coatings, which
makes spectrum interpretation easier. Sample
pump-down is even quicker.

A lot of people are going with this new generation of
SEMs. You still have the option of imaging the
regular way with these instruments.

My application is mostly material science. Being
concentrated in biology, I think you have a stronger
argument for buying an LV SEM. It would see a lot of
use in other disciplines once other people learn of
the instrument's capabilities.

Stu Smalinskas
Metallurgist
SKF
Plymouth, Michigan

__________________________________________________
Do you Yahoo!?
New DSL Internet Access from SBC & Yahoo!
http://sbc.yahoo.com


From daemon Wed Oct 2 11:24:18 2002



From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Wed, 02 Oct 2002 09:15:53 -0700
Subject: Re: Ask-A-Microscopist:ESEM Low Vacuum Mode. how useful is it?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Michael,
We recently purchased a new SEM and got the variable-pressure option, since
all the manufacturers offer this now and it doesn't cost much more. Ours is
a materials engineering department, but the microscopes are used by all
sorts of university and commercial users, including biologists.
I was pleasantly surprised with how useful the VP is for materials
applications. Samples that are not high vacuum compatible, such as wet or
oily stuff, samples that are impossible to coat or that cannot be coated
because they are needed in their current state for something else or just
things you want to look at very quickly, all work fine in VP mode, either by
back-scattered electron mode or a special secondary electron detector for
variable pressure. Apparently, with a cool or cold stage, the applications
are extended even more. That is next on my wish list. I would not consider
buying a conventional SEM without variable pressure option, just for the
versatility it offers.
I have no experience with the FEI ESEM, which goes to a much higher pressure
than the variable pressure range of my microscope.
At 12:53 AM 10/02/2002 -0500, you wrote:
}
} Email: mgrace-at-fit.edu
} Name: Michael Grace
}
} Organization: Florida Institute of Technology
}
} Education: Graduate College
}
} Location: Melbourne, FL, USA
}
} Question: I am working toward aquisition of a new SEM that will see
} use mostly in biological sciences applications, but will also see
} some use from materials scientists, chemists, etc. I am considering
} one of the new ESEMs (specifically the FEI/Philips Quanta 200.
} Question: how useful is the low vacuum mode, esp. for biological
} applications? Also interested in feedback from those with other
} applications, but I mainly want to be sure that the money is spent
} wisely- that we are not getting all the bells and whistles simply
} because they exist.
}
} Thank you.
}
Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca



From daemon Wed Oct 2 12:02:37 2002



From: Dusevich, Vladimir :      dusevichv-at-umkc.edu
Date: Wed, 2 Oct 2002 11:53:51 -0500
Subject: RE: Ask-A-Microscopist: SEM: Cambridge Stereoscan questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Excellent point.
And just some additional recommendations.

SEM resolution is highly specimen dependant. If you do not
have standard specimens for checking resolution, you can try
(for magnifications up to 10k) fractured metal. Do not use
non-conducting coated specimens.

Use shortest possible working distance.

If you cannot get rid completely of astigmatism with
electronic controls you should clean and/or realign
column.

By the way, what is "a few thousand X magnification"? 10k
could be pretty good for 1978 vintage SEM.

Vladimir


} Yes, but are you experienced in electron microscopes? This
} is definitely a
} field that not only encompasses those mentioned, but uses
} them, as well as
} others, all in synergistic manner.
}
} When it comes to resolution, look to your friend, the
} condenser control (or
} spot size or beam current or whatever else it may be called).
} On that
} vintage of Cambridge, I expect it is called spot size. If
} everything else
} is OK, the condenser is the primary control that determines
} the resolution
} of the instrument.
}
} This is because the primary aberration in an electron optic
} instrument is
} chromatic aberration. That is the result of the wide spread
} in energy
} ranges of the electrons emitted from the electron gun. That
} spread creates
} what is known as a spherical aberration that affects the beam
} in both the X
} and Y directions. As they travel through the various electromagnetic
} lenses of the SEM, electrons of different energies are
} affected differently
} primarily resulting in a different focal point in the Z axis. The
} condenser lens, and associated apertures, are used to reduce
} this energy
} spread.
}
} Here's the deal. The higher the condenser current, the
} smaller the spot
} size and the smaller the beam current. It's hard to explain
} in text, but
} you actually have two condenser lenses. Their purpose is to
} provide a
} de-magnification (reduction in apparent size) of the source of the
} electrons. The source is the tungsten filament, or cathode,
} that emits
} electrons from a rather large area of around 100 microns
} (micrometers).
}
} In the process, there are a couple of imaginary images
} formed, or crossover
} points, where apertures or electron absorbers with small
} central openings
} are located. Electrons of different energies will be
} affected differently
} by the electromagnetic lenses - electrons of lower energy
} will be displaced
} more than those of higher energies. The apertures will capture those
} electrons that have an energy lower or higher than that which
} the lenses
} focus through the apertures, thus reducing the beam current
} by removing
} those electrons with energies outside of the nominal and
} reducing the spot
} size by ensuring that those electrons remaining have a
} smaller spread of
} energies.
}
} Having said that, increase the condenser current, reduce the
} spot size or
} reduce the beam current, whichever applies.
}
} If you're new to electron microscopy, there is one thing that
} really must
} be stressed - cleanliness. The smallest, microscopic, piece
} of dust, lint,
} fiber or electrically insulating contamination can wreak
} havoc inside an
} electron optics column. They tend to build up a negative
} electrical charge
} under influence of the beam that can create repulsive fields
} that deflect
} the electron beam and create weird aberrations. Contamination from
} fingerprints and solvents is extremely minor compared to the
} effects of a
} single fiber in the electron path.
}
} Lastly, what ultimately determines the resolution possible in
} an instrument
} is the care with which it was installed. That includes the
} siting of the
} instrument in regards to mechanical vibrations in the
} building as well as
} the sources of electromagnetic interference. But most
} important is the
} careful routing of the various wires and tubing within and
} outside the
} instrument. The electron optics column is mounted on a table that is
} isolated from vibrations from the frame that holds it. All
} connections
} that run to this table must be carefully routed in order to
} provide similar
} isolation. Of similar concern is the electrical grounding of
} the various
} components of an SEM. Do not assume that because various large metal
} components appear to have electrical continuity that they do.
} Manufacturers typically provide for individual grounds to
} many components
} for good reason.
}
}
}
}
} Allen R. Sampson
} Advanced Research Systems
} 317 North 4th. Street
} St. Charles, Illinois 60174
}
} phone (630) 513-7093 fax (630) 513-7092 http://www.sem.com
}
}
} On Tuesday, October 01, 2002 10:51 PM, by way of MicroscopyListserver
} [SMTP:little-at-earthtech.org] wrote:
} }
} --------------------------------------------------------------
} ----------
} } The Microscopy ListServer -- Sponsor: The Microscopy
} Society of America
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} --------------------------------------------------------------
} ---------.
} }
} }
} } Below is the result of your feedback form (NJZFM-ultra-55). It was
} } submitted by (little-at-earthtech.org) from
} } http://www.msa.microscopy.com/Ask-A-Microscopist.html on Tuesday,
} } October 1, 2002 at 21:27:21
} }
} --------------------------------------------------------------
} -------------
} }
} } Email: little-at-earthtech.org
} } Name: George Luce and Scott Little
} }
} } Organization: Earthtech International, Inc.
} }
} } Education: Graduate College
} }
} } Location: Austin, Texas
} }
} } Question: We have recently revived a Cambridge Stereoscan 150 SEM
} } from about 1978. The machine is now operational, but we cannot get
} } good resolution or focus above a few thousand X magnification.
} }
} } We would like to communicate with someone familiar with this or a
} } similar model, who could give us some guidance on improving the
} } resolution. Details of our efforts can be supplied. We are
} } technically competent people with backgrounds in physics,
} } engineering, high voltage, electronics, vacuum, etc.
} }
} } Thank you.
} }
} }
} } George Luce (geolucetx-at-aol.com)
} }
} } Scott Little (little-at-earthtech.org)
} }
} }
} } Earthtech International, Inc.
} } Austin, Texas
} }
} } Phone (512) 342-2185
} }
} } www.earthtech.org
} }
} }
} --------------------------------------------------------------
} -------------
} }
}
}
}


From daemon Wed Oct 2 13:42:36 2002



From: sghoshro-at-NMSU.Edu
Date: Wed, 2 Oct 2002 12:31:22 -0600 (MDT)
Subject: Re: Ask-A-Microscopist:ESEM Low Vacuum Mode. how useful is it?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Michael,

The low vacuum mode is extremely useful for biological samples. For low
magnification pictures it is extremely useful and these days we hardly use
our sputter coater and critical point dryer, no fixation (you avoid use of
hazardous fixatives in that way). We are very satisfied with the results.
We have a Hitachi 3200N variable pressure SEM.

Good luck,

Soumitra Ghoshroy


*************************************************************
Soumitra Ghoshroy Ph.D.
Director, Electron Microscopy Lab
Graduate Faculty Member, Department of Biology
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
URL:http://confocal.nmsu.edu/eml

On Wed, 2 Oct 2002, by way of MicroscopyListserver wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (mgrace-at-fit.edu) from
} http://www.msa.microscopy.com/Ask-A-Microscopist.html on Tuesday,
} October 1, 2002 at 15:16:59
} ---------------------------------------------------------------------------
}
} Email: mgrace-at-fit.edu
} Name: Michael Grace
}
} Organization: Florida Institute of Technology
}
} Education: Graduate College
}
} Location: Melbourne, FL, USA
}
} Question: I am working toward aquisition of a new SEM that will see
} use mostly in biological sciences applications, but will also see
} some use from materials scientists, chemists, etc. I am considering
} one of the new ESEMs (specifically the FEI/Philips Quanta 200.
} Question: how useful is the low vacuum mode, esp. for biological
} applications? Also interested in feedback from those with other
} applications, but I mainly want to be sure that the money is spent
} wisely- that we are not getting all the bells and whistles simply
} because they exist.
}
} Thank you.
}
} ---------------------------------------------------------------------------
}
}



From daemon Wed Oct 2 15:17:11 2002



From: Dusevich, Vladimir :      dusevichv-at-umkc.edu
Date: Wed, 2 Oct 2002 15:07:48 -0500
Subject: RE: Ask-A-Microscopist:ESEM Low Vacuum Mode. how useful is it?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


FEI's ESEMs are equipped with good gaseous secondary electron
detectors and for "good" specimens their resolution in wet
mode is pretty close to the resolution in conventional mode.
They are widely used in materials science.

Nevertheless, in biological applications most specimens are
not "good". The best results I have had with hard tissue,
dentin and bone. In wet mode with wet dentin or bone I can
use magnification up to 25k, but if these specimens are dry
then in the same conditions (wet mode) magnification is
limited to 10k. With Au-Pd coated specimens in high vacuum
mode I can use magnifications up to 150k (I have a field
emission ESEM). I have had good results in observation
plants and (surprise!) bugs. Success with cell cultures
was marginal. And I have to mention that application of
EDS analysis in wet mode could be complicated and limited.

However, with all these limitations, I have found that
ESEM is a useful tool in our research.

Vladimir M. Dusevich, Ph.D.
Electron Microscope Lab Manager
3127 School of Dentistry
650 E. 25th Street
Kansas City, MO 64108-2784

Phone: (816) 235-2072
Fax: (816) 235-5524
Web: http://www.umkc.edu/dentistry/microscopy


} Email: mgrace-at-fit.edu
} Name: Michael Grace
}
} Organization: Florida Institute of Technology
}
} Education: Graduate College
}
} Location: Melbourne, FL, USA
}
} Question: I am working toward aquisition of a new SEM that will see
} use mostly in biological sciences applications, but will also see
} some use from materials scientists, chemists, etc. I am considering
} one of the new ESEMs (specifically the FEI/Philips Quanta 200.
} Question: how useful is the low vacuum mode, esp. for biological
} applications? Also interested in feedback from those with other
} applications, but I mainly want to be sure that the money is spent
} wisely- that we are not getting all the bells and whistles simply
} because they exist.
}
} Thank you.
}
} --------------------------------------------------------------
} -------------
}
}


From daemon Wed Oct 2 15:20:24 2002



From: Gerroir, Paul J :      Paul.Gerroir-at-crt.xerox.com
Date: Wed, 2 Oct 2002 15:56:47 -0400
Subject: TEM-SEM: Surplus Kevex Components

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Good Afternoon,

Following is a description of surplus Kevex equipment. If you are interested
in any of these items and are willing to pay shipping expenses please
contact me.

Regards,
Paul

Paul J. Gerroir
Microscopy
Materials Characterization
Xerox Research Centre of Canada
2660 Speakman Drive
Mississauga, Ontario L5K 2L1

Phone: (905) 823-7091, ext. 216
FAX: (905) 822-7022
email: paul.gerroir-at-crt.xerox.com


Description of Used Kevex Components


1. Kevex DeltaPlus Microanalysis System, Level 5 (without detector)
Serial #5500240-0158
Installed: November 1992
Electronics cabinet houses the signal processing electronics, MCA,
12-slot
dual wide DEC Q-bus card cage with LSI-11/73 minicomputer, dedicated
microprocessor and associated logic circuitry and data storage memory.
Kevex model 4461A pulse processor. Single Syquist drive available.

2. Kevex Delta Analyzer, complete system, however, light element detector
has 'blown' window
Serial #501678-1172
Detector: 30mm square, model 3600-0674
Serial #0790-5803
Pre-amplifier model 2003
Installed: October 1990




From daemon Wed Oct 2 16:48:47 2002



From: John Twilley :      jtwilley-at-sprynet.com
Date: Wed, 02 Oct 2002 17:51:31 -0400
Subject: Re: Wood Identification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Chuck,

To be very precise, crushed slivers may not be adequate, since some
distinctions are based on the variations in cell shape that occur through an
annual growth ring. You might want to contact Arno Schniewind :

arnops-at-nature.berkeley.edu

of the Forest Products Laboratory of the University of California, or his
colleagues:

http://www.ucfpl.ucop.edu/

Arno has been very helpful over the years and the lab provides wood
identification services.

John Twilley
Conservation Scientist

Chuck.Butterick-at-degussa.com-at-sparc5.microscopy.com wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Listers,
}
} Would someone please direct me to some resource, a good text, or better
} yet, a web site that can help me identify wood types by either cross-grain
} and/or with grain using a microscope? I have some crushed slivers that I'd
} like to identify. Any help would be appreciated. Thanks in advance.
}
} Chuck Butterick
} Degussa Engineered Carbons
} Borger, TX





From daemon Wed Oct 2 18:11:23 2002



From: Yan Xin :      xin-at-magnet.fsu.edu
Date: Wed, 02 Oct 2002 19:01:11 -0400
Subject: Thanks for the replies concerning EDS in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,

I would like to thank all the people for their helpful suggestions and
knowledge concerning the quantitative composition analysis in a TEM.

I did not give the details clearly when I posted the question.
I'd just like to provide more information on the detector and the
microscope I used.
The microscope is a Jeol-2010 LaB6 (though I wish it was a FEG) operated at
200kV.
The detector is a PGT Ge detector with a ultra thin window. The software is
called imix.
The atomic fraction I got for the EuS is calculated automatically from this
software after putting in voltage, take-off angle, density, and sample
thickness.

Just for your information.

Best Regards
Yan Xin





From daemon Wed Oct 2 20:57:04 2002



From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Wed, 02 Oct 2002 09:15:53 -0700
Subject: Re: Ask-A-Microscopist:ESEM Low Vacuum Mode. how useful is it?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Dear Michael,
We recently purchased a new SEM and got the variable-pressure option, since
all the manufacturers offer this now and it doesn't cost much more. Ours is
a materials engineering department, but the microscopes are used by all
sorts of university and commercial users, including biologists.
I was pleasantly surprised with how useful the VP is for materials
applications. Samples that are not high vacuum compatible, such as wet or
oily stuff, samples that are impossible to coat or that cannot be coated
because they are needed in their current state for something else or just
things you want to look at very quickly, all work fine in VP mode, either by
back-scattered electron mode or a special secondary electron detector for
variable pressure. Apparently, with a cool or cold stage, the applications
are extended even more. That is next on my wish list. I would not consider
buying a conventional SEM without variable pressure option, just for the
versatility it offers.
I have no experience with the FEI ESEM, which goes to a much higher pressure
than the variable pressure range of my microscope.
At 12:53 AM 10/02/2002 -0500, you wrote:
}
} Email: mgrace-at-fit.edu
} Name: Michael Grace
}
} Organization: Florida Institute of Technology
}
} Education: Graduate College
}
} Location: Melbourne, FL, USA
}
} Question: I am working toward aquisition of a new SEM that will see
} use mostly in biological sciences applications, but will also see
} some use from materials scientists, chemists, etc. I am considering
} one of the new ESEMs (specifically the FEI/Philips Quanta 200.
} Question: how useful is the low vacuum mode, esp. for biological
} applications? Also interested in feedback from those with other
} applications, but I mainly want to be sure that the money is spent
} wisely- that we are not getting all the bells and whistles simply
} because they exist.
}
} Thank you.
}
Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca




From daemon Thu Oct 3 07:51:50 2002



From: David Rothbard :      rothbardD-at-netscape.net
Date: Thu, 03 Oct 2002 08:37:39 -0400
Subject: Re: wood identification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Here are some texts on wood identification:

Hoadley, R.B. (1990) Identifying Wood: Accurate Results with Simple
Tools. Taunton Books. 1990 (try Amazon.com)
Parham, R. and Gray, R. (1982) The Practical Identification of Wood Pulp
Fibers. TAPPI Press, 212 pp. 1982.(try www.tappi.org)
Strelis, I. and Kennedy, R.W. (1967) Identification of North American
Commercial Pulpwoods and Pulp Fibers. University of Toronto Press. 1967
Core, H.A., Cote, W.A., Day, A. C. (1979) Wood Structure and
Identification. 1979


If you only have slivers, it can be difficult, expecially with
conifers. I can recommend some experienced analysts off-line.


David Rothbard
Earthborn Solutions
rothbardD-at-netscape.net


}
} Would someone please direct me to some resource, a good text, or better
} yet, a web site that can help me identify wood types by either
cross-grain
} and/or with grain using a microscope? I have some crushed slivers
that I'd
} like to identify. Any help would be appreciated. Thanks in advance.
}
} Chuck Butterick
} Degussa Engineered Carbons
} Borger, TX





From daemon Thu Oct 3 07:51:55 2002



From: Gerroir, Paul J :      Paul.Gerroir-at-crt.xerox.com
Date: Thu, 3 Oct 2002 08:24:35 -0400
Subject: LM: Liquid Crystals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Good morning all,

Could someone recommend a book, journal article or reference on the topic of
liquid crystals that encompasses both an introduction to the topic and
techniques for their characterization by light microscopy. Thanks.

Cheers,
Paul

Paul J. Gerroir
Microscopy
Materials Characterization
Xerox Research Centre of Canada
2660 Speakman Drive
Mississauga, Ontario L5K 2L1

Phone: (905) 823-7091, ext. 216
FAX: (905) 822-7022
email: paul.gerroir-at-crt.xerox.com



From daemon Thu Oct 3 10:26:41 2002



From: Marilyn Levy :      mlevy-at-cellbio.wustl.edu
Date: Thu, 3 Oct 2002 10:10:21 -0500
Subject: TEM(plastic embedding of tendon)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I have been asked to process adult mouse tendon, a new tissue for me. I
would appreciate any protocols or suggestions for processing this tissue
from fixation through TEM plastic embedding. Also, should I be concerned
about damage to my diamond knife during thin sectioning? Thanks in advance.
Marilyn Levy


From daemon Thu Oct 3 10:26:46 2002



From: Bruce Girrell :      bigirrell-at-microlinetc.com
Date: Thu, 3 Oct 2002 11:01:53 -0700
Subject: RE: Wood Identification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


} Would someone please direct me to some resource, a good text, or better
} yet, a web site that can help me identify wood types by either cross-grain
} and/or with grain using a microscope?

R. Bruce Hoadley is an expert on wood identification. If your sample size is
relatively large, i.e., splinter sized, rather than completely crushed, then
you can probably get pretty far just from Hoadley's book _Identifying Wood_
which is still in print (Taunton Press) - about US$30 from Amazon.com. If
not, you might try contacting Dr. Hoadley directly to see what he suggests.

Bruce Girrell






From daemon Thu Oct 3 13:59:24 2002



From: Sales Department :      sales-at-smoking.com.cn
Date: Thu, 3 Oct 2002 08:42:33 +0200
Subject: Discounted Cigs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sir or Madam

In the past you have requested information on discounted products. If you are not a smoker, and find this email offensive, then we sincerely apologise. We will be only too happy to take you off our database.

If you are a smoker, however, you are probably fed up with paying high prices for your cigarettes and tobacco. Take a look at what we can do for you at
http://www.smokersassociation.co.uk/?S=32&ID=2

We can send you, legally, by registered air mail, direct to your door, 4 cartons of cigarettes or 40 pouches of rolling tobacco (all brands are available) from only 170 Euros - about 105 pounds - fully inclusive of postage and packing. Why pay more?

If you would rather not hear from us any more, this link will ensure that you are not bothered again.
mailto:unsubscribe-at-smokersassociation.co.uk

Yours faithfully.
Smokers Association

http://www.smokersassociation.co.uk/?S=32&ID=2

xay7712041y

From daemon Thu Oct 3 15:11:14 2002



From: Erik Bodegom :      Bodegom-at-pdx.edu
Date: Thu, 03 Oct 2002 13:02:26 -0700
Subject: position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Microscopist and Lab Manager for a Centralized Electron Microscope
Laboratory Facility at Portland State University

Portland State University seeks an electron microscopist to operate and
manage a newly established electron microscopy (EM) facility consisting of a
FEI/Philips (Tecnai F-20) 200kV field emission high-resolution transmission
electron microscope (TEM) equipped with an embedded digital scanning
transmission electron microscopy (STEM) capability, and energy dispersive
x-ray spectrometer (EDS), a JEOL 2000FX TEM, and a 611 FEI focused ion beam
microscope.

PSU is an AA/EO institution and, in keeping with the president's diversity
imitative, welcomes applications from diverse candidates and candidates who
support diversity.

The responsibilities of the position include the management, operation, and
maintenance of the microscopes, and training and assisting faculty, and
student, and outside university users of the microscopes.

Candidates preferentially have at least 5-years experience in managing an
electron microscopy multi-user facility in an academic or industrial
research setting and an outstanding record of team and personal
accomplishments. A publication record of EM-related research is expected.
Degree in electron microscopy and relevant disciplines in using microscopes
is required.Advanced university degree is required as well as a demonstrated
experiences in either materials science, geological, and or biological
materials characterizations are essential. The candidates must have working
knowledge onexperience using the microscopes listed above and are be
familiar such operational techniques as digital image acquisition, image
processing (digital micrograph) , structure simulation, and EDS analysis.

The candidates should also be able to have experience performing minor
microscope repairs and maintenance. repair minor problems of the
microscopes while mMicroscopes will be maintained under ajor components of
the microscopes will be covered by the service contracts.

The successful candidates are expected to have be self-motivationmotivated,
patience, and the ability to work and work well with exiting technical
staff, faculty, and students, and outside users.

Candidates will be selected based on their qualifications and
accomplishments. Salary will be commensurate with qualifications and
experience.

Review of applications will begin on October 1, 2002, and continue until the
position is filled.

Interested candidates, please should send a letter of interest with a
two-page management plan for a multi-user EM facility and a full resume and
arrange to have five letters of reference, sent to the address below:

Search Committee for EM Manager
Physics DepartmentDepartment of Physics
Portland State University
P.O. Box 751
Portland, OR 97207-0751





From daemon Thu Oct 3 22:57:51 2002



From: Sales Department :      sales-at-smoking.com.cn
Date: Thu, 3 Oct 2002 08:42:33 +0200
Subject: Discounted Cigs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Dear Sir or Madam

In the past you have requested information on discounted products. If you are not a smoker, and find this email offensive, then we sincerely apologise. We will be only too happy to take you off our database.

If you are a smoker, however, you are probably fed up with paying high prices for your cigarettes and tobacco. Take a look at what we can do for you at
http://www.smokersassociation.co.uk/?S=32&ID=2

We can send you, legally, by registered air mail, direct to your door, 4 cartons of cigarettes or 40 pouches of rolling tobacco (all brands are available) from only 170 Euros - about 105 pounds - fully inclusive of postage and packing. Why pay more?

If you would rather not hear from us any more, this link will ensure that you are not bothered again.
mailto:unsubscribe-at-smokersassociation.co.uk

Yours faithfully.
Smokers Association

http://www.smokersassociation.co.uk/?S=32&ID=2

xay7712041y

From daemon Fri Oct 4 09:17:23 2002



From: Scott Whittaker :      Whittaker.scott-at-nmnh.si.edu
Date: Fri, 04 Oct 2002 10:05:31 -0400
Subject: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I have been coating samples via electron beam evaporation to a thickness of
20-25 nm as determined by blue interference colors on polished brass.
Recently the mineral guys lost their coater so I have also been coating
their material to the same blue color they said they used as well. They used
resistance heating.

In examining those samples they noticed their EDS/WDS results appearing a
little low so we went ahead and coated the standards as well.

More digging has revealed they coat with graphite rods whereas I have been
using amorphous C rods. They swear the coats are thicker and looking at
theirs-vs-mine it appears so BUT we both evaporated to the same blue color
on brass.

Could there be a difference in the optical properties of the graphite and
amorphous rods?? I assumed maybe incorrectly that the evaporated material
was basically the same from the two rods thus blue brass would also equal
the same thickness. Thanks

Scott Whittaker
Lab Manager
Laboratories of Analytical Biology
Smithsonian Institution
National Museum of Natural History
PO Box 37012 MRC104
Washington DC 20013-7012
202-357-1651



From daemon Fri Oct 4 09:50:44 2002



From: Barbara Maloney :      maloneyb-at-fiu.edu
Date: Fri, 04 Oct 2002 10:39:36 -0400
Subject: PanaCL and/or cold stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi group - anybody out there currently using a CL detector with a cold
stage - please email me ASAP
Thanks so much
Barb



From daemon Fri Oct 4 10:07:48 2002



From: S. Kuehner :      kuehner-at-u.washington.edu
Date: Fri, 4 Oct 2002 08:00:34 -0700 (PDT)
Subject: Re: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



I hope someone knows the answer to this question because I've wondered the
same thing... carbon rods vs graphite. another scott

************************************************
....amphiboles do violence to history...
T. Feininger, 2001. (taken out of context)
****************************

Dr. Scott Kuehner kuehner-at-u.washington.edu
Dept. of Geological Sciences ph.206-543-8393
Box 351310 Fax 206-616-6873
The University of Washington
Seattle, Washington 98195-1310
************************************************

On Fri, 4 Oct 2002, Scott Whittaker wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I have been coating samples via electron beam evaporation to a thickness of
} 20-25 nm as determined by blue interference colors on polished brass.
} Recently the mineral guys lost their coater so I have also been coating
} their material to the same blue color they said they used as well. They used
} resistance heating.
}
} In examining those samples they noticed their EDS/WDS results appearing a
} little low so we went ahead and coated the standards as well.
}
} More digging has revealed they coat with graphite rods whereas I have been
} using amorphous C rods. They swear the coats are thicker and looking at
} theirs-vs-mine it appears so BUT we both evaporated to the same blue color
} on brass.
}
} Could there be a difference in the optical properties of the graphite and
} amorphous rods?? I assumed maybe incorrectly that the evaporated material
} was basically the same from the two rods thus blue brass would also equal
} the same thickness. Thanks
}
} Scott Whittaker
} Lab Manager
} Laboratories of Analytical Biology
} Smithsonian Institution
} National Museum of Natural History
} PO Box 37012 MRC104
} Washington DC 20013-7012
} 202-357-1651
}
}
}



From daemon Fri Oct 4 10:44:40 2002



From: Jiao, Chengge :      chengge.jiao-at-uk.feico.com
Date: Fri, 4 Oct 2002 17:29:21 +0200
Subject: A SIMS book

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

Does anybody know where I can buy a book called "Secondary Ion Mass Spectrometry: Basic concepts, Instrumental aspects, applications and trends " by Benninghoven, A.; F.G.Rudenauer F.G., and Werner H.W.

Thanks for your information.

Chengge Jiao


From daemon Fri Oct 4 14:58:23 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Fri, 04 Oct 2002 13:11:11 -0700
Subject: Re: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I used to use carbon/graphite rods for making thin carbon film. I do find
that using thermal evaporation, the properties of the film originated from
graphite or "carbon" are different. It looks like, "graphite" films are
more fragile, less stable and somehow more "crumbly". I do notice that
"graphite" evaporates at lower current (thermal) also. This difference (in
properties of the film) is less pronounced when Electron Beam Evaporator
(Electron Gun) was used... But, still, I prefer to use spectrographic
quality "carbon rods".

In general, carbon thickness is difficult to determine precisely. I am
using Maxtec TM-450(?) thickness monitor with preciseness 0.01 nm. Even
with this instrument, the real thickness may vary up to 10-15% (never
measured exact value, but have "feelings"). Sergey


At 08:00 AM 10/4/02 -0700, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

------------------------------------------------------

Sergey Ryazantsev, Ph.D.
Electron Microscopy
Department of Biological Chemistry, School of Medicine
University of California, Los Angeles
Box 951737
Los Angeles, CA 90095-1737

(310) 825-1144 (office)
Pager: (310) 845-0248
FAX: (310) 206-5272 (departmental)
mailto:sryazant-at-ucla.edu



From daemon Fri Oct 4 14:58:23 2002



From: Mark Aindow :      m.aindow-at-uconn.edu
Date: Fri, 4 Oct 2002 15:51:28 -0400
Subject: Postdoctoral Position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



University of Connecticut
Institute of Materials Science

Postdoctoral Position
Displacive Phase Transformations

A post-doctoral position is available immediately within the Institute
of Materials Science at the University of Connecticut to work on
displacive phase transformations in crystalline solids. The systems to
be studied include metallic pseudoelastic alloys and ferroelectric thin
films. The ideal candidate will have experience in both
microstructural characterization (TEM, SEM etc.) and
phenomenological/crystallographic modeling of phase transformations.
This position is a one-year appointment, with possible renewal for a
second year.

To apply, please send a complete resume, together with a list of
publications and contact details for 3 referees to either Prof. S.
Pamir Alpay (p.alpay-at-ims.uconn.edu) or Prof. M. Aindow
(m.aindow-at-uconn.edu). Screening of applications will begin immediately,
and will continue until the position is filled. We encourage
applications from under-represented groups, including minorities, women
and people with disabilities.


*********************************************************

Mark Aindow, Associate Professor,
Department of Metallurgy and Materials Engineering
Institute of Materials Science,
97 North Eagleville Road, Unit 3136
University of Connecticut, Storrs,
CT 06269-3136, USA

Tel: +1 (860) 486-2644
FAX: +1 (860) 486-4745
Email: m.aindow-at-uconn.edu

**********************************************************



From daemon Fri Oct 4 15:02:12 2002



From: John J. Bozzola :      bozzola-at-siu.edu
Date: Fri, 4 Oct 2002 14:56:06 -0500
Subject: CSMMS Meeting Announcement (Call for Papers)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


FIRST CALL FOR PRESENTATIONS & FALL MEETING ANNOUNCEMENT

SOCIETY: Central States Microscopy and Microanalysis Society (CSMMS)

DATE: November 15, 2002. From 9 AM to 4 PM.

LOCATION: Giant City State Park. Carbondale, Illinois

DESCRIPTION:

The CSMMS will hold its Fall meeting at the rustic lodge at the Giant
City State Park in Carbondale, Illinois. Located in a lovely wooded
park, the lodge will serve as the focal point for the presentations
and luncheon banquet. The registration and orientation will begin at
9 AM and presentations will commence at 9:30 AM.

The first half-day will involve research presentations, followed by a
luncheon and business meeting. The second half-day will involve
technical presentations, demonstrations and discussions. For example,
we will have presentations on digital imaging (Production of Large
Format Posters, Digital Publishing, etc.) as well as discussions of
useful techniques and technical pointers. Topics may cover any type
of microscopy and microanalysis.

If you are interested in attending or presenting, please contact me
for details.

##############################################################
John J. Bozzola, Ph.D., Director
I.M.A.G.E. (Integrated Microscopy & Graphics Expertise)
750 Communications Drive - MC 4402
Southern Illinois University
Carbondale, IL 62901 U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
##############################################################









From daemon Fri Oct 4 20:55:18 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Fri, 04 Oct 2002 18:47:26 -0700
Subject: Re: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Tobias, good question

In specification it indicates 0.01 nm accuracy as far as I remember, but as
I mentioned in my original posting, it seems to me that I do have
approximately 10% difference in the film thickness even with that smart
thickness monitor. In general, I am happy with that instrument. It's very
solid and sensitive enough in 10-20 A scale. Sergey

At 04:15 PM 10/4/02, you wrote:
} Dear Sergey,
} Does your thickness monitor really measure down to 0.01 nm
} accuracy?? That's amazing.
}
} Tobias
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Fri Oct 4 22:05:22 2002



From: Johnson Mulete :      john_2007-at-cheerful.com
Date: Sat, 05 Oct 2002 04:56:06
Subject: Private Proposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Re: Business Proposal

Sir/Madam,

I am JOHNSON MULETE JR. the son of Mr. Luke Radebe MULETE from Zimbabwe. I am sorry this mail
Will Surprise you, though we do not know, I got your contact Through South African Chamber of Commerce.

During the current war against white farmers in Zimbabwe and the support of President Robert Mugabe to
Claim all the white owned farms in our country to gain Favor for re-election. All white farmers were asked to
Surrender their farms to the government for Re-distribution and infact to his political party
Members and my father though black, was the treasury Of the farmers association and a strong member of an
Opposition party that did not support the presidents Idea. He then ordered his party members and the police
Under his pay role to invade my fathers farm and burn Down everything in the farm. They killed my
Father and took away a lot of items from his farm.

After the death of my father, our local pastor and a Close friend of my father handed us over will
Documents with instructions from my father that we Should leave Zimbabwe for South Africa incase any
Thing happen to him, process the release of the fund Kept in custody for us in a security company unknown
To the company that the content is money hence it was deposited as Personal Belongings and Ensure that we do not remain there as we could Easily be found by his enemies. The total amount is US$21.5M.We are therefore soliciting for
Your assistance to help us move the fund out of South Africa, as our fate and future is far from Reality, Hence this mail to you. The president's present ban of International Press into Zimbabwe is just a few of the Unthinkable things he is committing in my Country.

I have tried to reach my father's close friend Mr. John Casabas from Australia also a farmer who was
Leaving in Zimbabwe with us but left with his family Late last year following this ugly development to no Avail.

Should you be interested to help us, contact me Immediately via email for easy communication and I Will furnish you with the time frame and modalities of The transaction. Please note that this transaction is} 100% confidential and risk free and will not endanger You or us in any way. We have resolved to give you 20% Of the total sum upon confirmation of the fund in any Account of your choice were the incident of taxation Will not take much tool on the money and we look Forward to coming over to your country to invest our Share and settle there. I will provide you my phone number
As soon as you indicate your willingness to assist us so that our conversation can be 100%confidential.

Please do not use the reply button, reply only to(johnmulte-at-netscape.net} Please take note.

God bless you indeed as you help yourself and us.

MR. Johnson Mulete Jr.



From daemon Sat Oct 5 09:26:30 2002



From: charles j day :      wa5ekh-at-juno.com
Date: Fri, 4 Oct 2002 09:03:28 -0500
Subject: Re: Image Analysis Short Course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


ANYBODY perfected the SEM "shallow and deep(bottom and no bottom-black)
hole area(morphometry) size analysis??
I'm still having trouble with thresholding, flatness, edge defintion and
mostly repeatability and accuracy.
+_20-50% isn't of much use. Manual edge detection is too slow and not
much more repeatable. I'd like to see at least +_5%.
jeffrey day/mesqite(o), texas


From daemon Sat Oct 5 11:00:26 2002



From: Cle02JS12-at-aol.com (by way of Ask-A-Microscopist)
Date: Sat, 5 Oct 2002 10:47:38 -0500
Subject: Ask-A-Microscopist:LM American Optical ? Microscope repair parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (Cle02JS12-at-aol.com) from on Saturday, October 5, 2002
at 01:33:09
---------------------------------------------------------------------------

Email: Cle02JS12-at-aol.com
Name: Sandy

Education: 9-12th Grade High School

Location: Washington State

Question: Need referral on how/where to find replacement part for an
older college level microscope. Emblem indicates "Alpha
Omega-Spencer".
Under side indicates American Optical Co.

Any leads much appreciated.

Thank You



---------------------------------------------------------------------------


From daemon Sat Oct 5 15:21:04 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Sat, 5 Oct 2002 14:03:59 -0600
Subject: Re: Image Analysis Short Course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Jeffrey,

I am not quite sure what the problem is that you are referring to, but you
may want to check out our ACT module (check our web site). It is designed to
find edges and do geometric measurements on those edges, using tools
developed for the semiconductor industry. If you want to send me an image
for a trial run, please send it to my address below.

mike


Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: charles j day [mailto:wa5ekh-at-juno.com]
Sent: Friday, October 04, 2002 8:03 AM
To: DrJohnRuss-at-aol.com-at-sparc5.microscopy.com
Cc: microscopy-at-sparc5.microscopy.com; NIH-IMAGE-at-LIST.NIH.GOV


ANYBODY perfected the SEM "shallow and deep(bottom and no bottom-black)
hole area(morphometry) size analysis??
I'm still having trouble with thresholding, flatness, edge defintion and
mostly repeatability and accuracy.
+_20-50% isn't of much use. Manual edge detection is too slow and not
much more repeatable. I'd like to see at least +_5%.
jeffrey
day/mesqite(o), texas


From daemon Sat Oct 5 15:21:04 2002



From: Jim Quinn :      jquinn-at-www.matscieng.sunysb.edu
Date: Sat, 5 Oct 2002 16:10:40 -0400
Subject: Re: A SIMS book

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Cheng-Ge Jiao

try "www.google.com"
enter "Benninghoven Rudenauer"
brings up this URL:
http://www.addall.com/Browse/Detail/0471010561.html
states:

Secondary Ion Mass Spectrometry Basic Concepts, Instrumental Aspects, Applications & Trends
A. Benninghoven , H. W. Werner , F. G. Rudenauer
Binding: Hardcover, 1 edition, 1264 pages
Publisher: Wiley, John and Sons, Incorporated
Published Date: 01/01/1987
List: USD $395.00
ISBN: 0471010561


"click on" next link for list of locations:

http://www.addall.com/New/BrowseCompare.cgi?isbn=0471010561

good luck,

Jim

} From Microscopy-request-at-sparc5.microscopy.com Fri Oct 4 20:06:03 2002
} Subject: A SIMS book
} Date: Fri, 4 Oct 2002 17:29:21 +0200
} From: "Jiao, Chengge" {chengge.jiao-at-uk.feico.com}
} To: "Microscopy-at-MSA.Microscopy.Com" {Microscopy-at-sparc5.microscopy.com}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear All,
}
} Does anybody know where I can buy a book called
} "Secondary Ion Mass Spectrometry: Basic concepts,
} Instrumental aspects, applications and trends "
} by Benninghoven, A.; F.G.Rudenauer F.G., and Werner H.W.
}
} Thanks for your information.
}
} Chengge Jiao
}


From daemon Sun Oct 6 09:15:48 2002



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sun, 06 Oct 2002 10:06:14 -0500
Subject: Carbon vs. graphite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

There have been several postings on the subject of (real) carbon vs.
graphite rods for use in carbon coating in a vacuum evaporator.

In many instances, I find that our customers are purchasing "carbon" rods
but which are really graphite rods. I am not suggesting that those who
offer the product as being truly a carbon rod are doing something wrong
because it seems like when we use the words "carbon rods" we don't really
think of it as being carbon or graphite but somehow we know what we all mean
when we say we use "carbon rods".

But graphite (when used as a "carbon rod") is described in terms of its
porosity; indeed, there is a wide range of densities of graphite rods. It
is our perception, but not as a result of exhaustive testing, that the
higher the porosity ( and lower the density), when one evaporates, there is
increased "sparking". I think I would be safe in saying that the
"conventional wisdom" suggests that this is undesirable. At the wholesale
level (e.g. the large quantities that someone like SPI Supplies purchases)
there is in fact a substantial difference in the price between graphite rods
of higher porosity vs. lower porosity. We believe that the lower porosity
(and accompanying higher density) results in less sparking, perhaps less
contaminating nanotubes (which are a bi-product of this kind of evaporation)
and other artifactual features.

Actual "carbon" rods have the lowest porosity. Although we have not tested
it to the point that I would stake my life on it, real carbon rods seem to
spark a whole lot less (if at all) and the final film seems to be much more
featureless, but we think that is due to the reduced amount of "sparking".
Further information can be found at URL
http://www.2spi.com/catalog/spec_prep/carbon-graphite-rods.html

I would be most appreciative of any comments pro or con relative to this
analysis. I have been operating on the assumption that this is more or less
correct. If there are other points of view, I would be the first to want to
know what they are.

Disclaimer: SPI Supplies has offered for some time both carbon rods as
carbon rods and graphite rods as graphite rods.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================





-----------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America To
Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com On-Line
Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
-----------------------------------------------------------------------.


I used to use carbon/graphite rods for making thin carbon film. I do find
that using thermal evaporation, the properties of the film originated from
graphite or "carbon" are different. It looks like, "graphite" films are
more fragile, less stable and somehow more "crumbly". I do notice that
"graphite" evaporates at lower current (thermal) also. This difference (in
properties of the film) is less pronounced when Electron Beam Evaporator
(Electron Gun) was used... But, still, I prefer to use spectrographic
quality "carbon rods".

In general, carbon thickness is difficult to determine precisely. I am
using Maxtec TM-450(?) thickness monitor with preciseness 0.01 nm. Even
with this instrument, the real thickness may vary up to 10-15% (never
measured exact value, but have "feelings"). Sergey


At 08:00 AM 10/4/02 -0700, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

------------------------------------------------------

Sergey Ryazantsev, Ph.D.
Electron Microscopy
Department of Biological Chemistry, School of Medicine
University of California, Los Angeles
Box 951737
Los Angeles, CA 90095-1737

(310) 825-1144 (office)
Pager: (310) 845-0248
FAX: (310) 206-5272 (departmental)
mailto:sryazant-at-ucla.edu





From daemon Sun Oct 6 14:51:27 2002



From: charles j day :      wa5ekh-at-juno.com (by way of MicroscopyListserver)
Date: Sun, 6 Oct 2002 14:34:17 -0500
Subject: SEM Image Analysis-deep and shallow hole sizing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


ANYBODY perfected the SEM "shallow and deep(bottom and no bottom-black)
hole area(morphometry) size analysis??
I'm still having trouble with thresholding, flatness, edge defintion and
mostly repeatability and accuracy., etc.
+_20-50% isn't of much use. Manual edge detection is too slow and not
much more repeatable. I'd like to see at least +_5%.
jeffrey
day/mesqite(o), texas


From daemon Sun Oct 6 15:05:30 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Sun, 06 Oct 2002 13:22:20 -0700
Subject: Re: Carbon vs. graphite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Charles

Graphite and carbon rods I have (from Ted Pella, I believe) are easily
distinguished: graphite is grayish and carbon is deep black. There is
also difference in hardness: carbon is harder. I really like your point
that graphite produce more "features" when evaporated. It support my
intuitional decision to use "carbon". It seems to me it produces more
uniform film... Thanks. Sergey

At 10:06 AM 10/6/02 -0500, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Sun Oct 6 17:24:47 2002



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Mon, 07 Oct 2002 11:12:30 +1200
Subject: Re: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



} } } "crumbly". I do notice that "graphite" evaporates at lower current
} } } (thermal) also. This difference (in properties of the film) is less
} } } pronounced when Electron Beam Evaporator (Electron Gun) was used...
} } } But, still, I prefer to use spectrographic quality "carbon rods".
} } }


That's funny. When I inadvertantly changed over from 'carbon' to
'graphite, I found that the graphite needed so much current that the rod
holder on my Edwards 306 became very hot. I couldn't easily get
sufficient evaporation, even at maximum current, so I switched back to
carbon.

cheers

rtch

Ritchie Sims Phone : 64 9 3737599 ext 7713
Department of Geology Fax : 64 9 3737435
The University of Auckland email : r.sims-at-auckland.ac.nz
Private Bag 92019
Auckland
New Zealand



From daemon Sun Oct 6 19:45:05 2002



From: admin5217q00-at-easymoney.fr.com
Date: Mon, 07 Oct 2002 10:03:03 -1100
Subject: Welcome To the Club

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello!

You may have seen this business before and
ignored it. I know I did - many times! However,
please take a few moments to read this letter.
I was amazed when the profit potential of this
business finally sunk in... and it works!

With easy-to-use e-mail tools and opt-in e-mail,
success in this business is now fast, easy and
well within the capabilities of ordinary people
who know little about internet marketing. And the
earnings potential is truly staggering!

I'll make you a promise. READ THIS E-MAIL TO THE END! -
follow what it says to the letter - and you will not
worry whether a RECESSION is coming or not, who is
President, or whether you keep your current job or not.
Yes, I know what you are thinking. I never responded
to one of these before either. One day though,
something just said: You throw away $25.00 going to
a movie for 2 hours with your wife. What the heck.
Believe me, no matter where you believe those feelings
come from, I thank every day that I had that feeling.

I cannot imagine where I would be or what I would be
doing had I not. Read on. It's true. Every word of it.
It is legal. I checked. Simply because you are buying
and selling something of value.

AS SEEN ON NATIONAL TV:

Making over half a million dollars every 4 to 5
months from your home.

THANKS TO THE COMPUTER AGE AND THE INTERNET!
==================================================
BE AN INTERNET MILLIONAIRE LIKE OTHERS WITHIN A YEAR!!!

Before you say Bull, please read the following.
This is the letter you have been hearing about on the
news lately. Due to the popularity of this letter on
the internet, a national weekly news program recently
devoted an entire show to the investigation of this
program described below, to see if it really can make
people money. The show also investigated whether or
not the program was legal.

Their findings proved once and for all that there are
absolutely NO laws prohibiting the participation in
the program and if people can 'follow the simple
instructions' they are bound to make some mega bucks
with only $25 out of pocket cost.

DUE TO THE RECENT INCREASE OF POPULARITY & RESPECT
THIS PROGRAM HAS ATTAINED, IT IS CURRENTLY WORKING
BETTER THAN EVER.

This is what one had to say: Thanks to this
profitable opportunity. I was approached many times
before but each time I passed on it. I am so glad
I finally joined just to see what one could expect
in return for the minimal effort and money required.
To my astonishment, I received a total $610,470.00
in 21 weeks, with money still coming in.

Pam Hedland, Fort Lee, New Jersey.
==================================================
Another said: This program has been around for a
long time but I never believed in it. But one day
when I received this again in the mail I decided
to gamble my $25 on it. I followed the simple
instructions and walaa ..... 3 weeks later the
money started to come in. First month I only made
$240.00 but the next 2 months after that I made
a total of $290,000.00. So far, in the past 8 months
by re-entering the program, I have made over
$710,000.00 and I am playing it again. The key to
success in this program is to follow the simple
steps and NOT change anything.

More testimonials later but first, =======

==== PRINT THIS NOW FOR YOUR FUTURE REFERENCE ====
$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$

If you would like to make at least $500,000 every
4 to 5 months easily and comfortably, please read the
following...THEN READ IT AGAIN and AGAIN!!!

$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$

FOLLOW THE SIMPLE INSTRUCTION BELOW AND
YOUR FINANCIAL DREAMS WILL COME TRUE, GUARANTEED!

INSTRUCTIONS:

=====Order all 5 reports shown on the list below =====

For each report, send $5 CASH, THE NAME & NUMBER OF THE
REPORT YOU ARE ORDERING and YOUR E-MAIL ADDRESS to the
person whose name appears ON THAT LIST next to the report.
MAKE SURE YOUR RETURN ADDRESS IS ON YOUR ENVELOPE TOP
LEFT CORNER in case of any mail problems.

===WHEN YOU PLACE YOUR ORDER, MAKE SURE ===
===YOU ORDER EACH OF THE 5 REPORTS! ===
You will need all 5 reports so that you can save
them on your computer and resell them.
YOUR TOTAL COST $5 X 5 = $25.00.

Within a few days you will receive, via e-mail, each
of the 5 reports from these 5 different individuals.
Save them on your computer so they will be accessible
for you to send to the 1,000's of people who will
order them from you. Also make a floppy of these
reports and keep it on your desk in case something
happens to your computer.

IMPORTANT - DO NOT alter the names of the people who
are listed next to each report, or their sequence on
the list, in any way other than what is instructed
below in steps 1 through 6 or you will lose out
on the majority of your profits. Once you
understand the way this works, you will also
see how it will not work if you change it.

Remember, this method has been tested, and if you
alter it, it will NOT work!!! People have tried
to put their friends'/relatives' names on all five
thinking they could get all the money. But it does
not work this way. Believe us, some have tried to
be greedy and then nothing happened. So Do Not
try to change anything other than what is instructed.
Because if you do, it will not work for you.
Remember, honesty reaps the reward!!!

This IS a legitimate BUSINESS. You are offering a
product for sale and getting paid for it. Treat it
as such and you will be VERY profitable in a short
period of time.

1.. After you have ordered all 5 reports, take this
advertisement and REMOVE the name & address of the
person in REPORT # 5. This person has made it through
the cycle and is no doubt counting their fortune.

2.. Move the name & address in REPORT # 4 down TO REPORT # 5.

3.. Move the name & address in REPORT # 3 down TO REPORT # 4.

4.. Move the name & address in REPORT # 2 down TO REPORT # 3.

5.. Move the name & address in REPORT # 1 down TO REPORT # 2

6.... Insert YOUR name & address in the REPORT # 1 Position.

PLEASE MAKE SURE you copy every name & address
ACCURATELY! This is critical to YOUR success.

==================================================
**** Take this entire letter, with the modified
list of names, and save it on your computer.
DO NOT MAKE ANY OTHER CHANGES.

Save this on a disk as well just in case you lose
any data. To assist you with marketing your business
on the internet, the 5 reports you purchase will
provide you with invaluable marketing information
which includes how to send bulk e-mails legally,
where to find thousands of free classified ads and
much more. There are 2 primary methods to get this
venture going:

METHOD # 1: BY SENDING BULK E-MAIL LEGALLY
==================================================

Let's say that you decide to start small, just to see
how it goes, and we will assume you and those involved
send out only 5,000 e-mails each. Let's also assume
that the mailing receives only a 0.2% (2/10 of 1%)
response (the response could be much better but let's
just say it is only 0.2%). Also many people will send
out hundreds of thousands of e-mails instead of only
5,000 each.

Continuing with this example, you send out only
5,000 e-mails. With a 0.2% response, that is only
10 orders for report # 1. Those 10 people responded
by sending out 5,000 e-mails each for a total of
50,000. Out of those 50,000 e-mails only 0.2%
responded with orders. That's 100 people who
responded and ordered Report # 2.

Those 100 people mail out 5,000 e-mails each for
a total of 500,000 e-mails. The 0.2% response to
that is 1000 orders for Report # 3.

Those 1000 people send 5,000 e-mails each for a
total of 5 million e-mails sent out. The 0.2%
response is 10,000 orders for Report # 4.

Those 10,000 people send out 5,000 e-mails each
for a total of 50,000,000 (50 million) e-mails.
The 0.2% response to that is 100,000 orders for
Report # 5.

THAT'S 100,000 ORDERS TIMES $5 EACH = $500,000.00
(half a million dollars).

Your total income in this example is: 1..... $50
+ 2..... $500 + 3.....$5,000 + 4..... $50,000
+ 5.... $500,000 .... Grand Total=$555,550.00

NUMBERS DO NOT LIE. GET A PENCIL & PAPER AND
FIGURE OUT THE WORST POSSIBLE RESPONSES AND NO
MATTER HOW YOU CALCULATE IT, YOU WILL STILL
MAKE A LOT OF MONEY!
==================================================

REMEMBER FRIEND, THIS IS ASSUMING ONLY 10 PEOPLE
ORDERING OUT OF 5,000 YOU MAILED TO. Dare to think
for a moment what would happen if everyone or half
or even one 4th of those people mailed 100,000
e-mails each or more?

There are over 150 million people on the internet
worldwide and counting, with thousands more coming
online every day. Believe me, many people will do
just that, and more!

METHOD # 2: BY PLACING FREE ADS ON THE INTERNET
==================================================

Advertising on the net is very, very inexpensive and
there are hundreds of FREE places to advertise.
Placing a lot of free ads on the internet will
easily get a larger response. We strongly suggest
you start with Method # 1 and add METHOD # 2 as you
go along. For every $5 you receive, all you must
do is e-mail them the report they ordered. That's it.
Always provide same day service on all orders.

This will guarantee that the e-mail they send out,
with your name and address on it, will be prompt
because they cannot advertise until they receive
the report.

===========AVAILABLE REPORTS ====================
The reason for the cash is not because this is
illegal or somehow wrong. It is simply about time.
Time for checks or credit cards to be cleared or
approved, etc. Concealing it is simply so no one
can SEE there is money in the envelope and steal
it before it gets to you.

ORDER EACH REPORT BY ITS NUMBER & NAME ONLY.
Notes: Always send $5 cash (U.S. CURRENCY or EURO) for
each report. Checks NOT accepted. Make sure the
cash is concealed by wrapping it in at least
2 sheets of paper. On one of those sheets of
paper, write the NUMBER & the NAME of the report
you are ordering, YOUR E-MAIL ADDRESS and your
name and postal address.

PLACE YOUR ORDER FOR THESE REPORTS NOW :
==================================================
REPORT # 1: The Insider's Guide To Advertising
for Free On The Net
Order Report #1 from:

EARL PACE
P.O. BOX 371
SUCC. ST-MARTIN
LAVAL
H7V 3P6
CANADA
______________________________________________________
REPORT # 2: The Insider's Guide To Sending Bulk
Email On The Net
Order Report # 2 from:

Celine Tremblay
P.O.BOX 49
SUCC. Cartieville
2127 Marcel-Laurin
Saint-laurent
Canada
H4R2V7

______________________________________________________

REPORT # 3: Secret To Multilevel Marketing On The Net
Order Report # 3 from:

Corine Dandurand
Coreex 109
19 Avenue du marechal Foch
77508 CHELLES CEDEX
FRANCE

______________________________________________________

REPORT # 4: How To Become A Millionaire Using
MLM & The Net
Order Report # 4 from:

Jean Lafortune
107, rue Lauriston
75116 Paris
FRANCE
_______________________________________________________

REPORT # 5: How To Send Out One Million Emails
For Free
Order Report # 5 From:

Stuart Iles
9 Elm Way
Hartshill
NUNEATON
CV10 0XS
United Kingdom

______________________________________________________
$$$$$$$$$ YOUR SUCCESS GUIDELINES $$$$$$$$$$$

Follow these guidelines to guarantee your success:

=== If you do not receive at least 10 orders for
Report #1 within 2 weeks, continue sending e-mails
until you do.

=== After you have received 10 orders, 2 to 3 weeks
after that you should receive 100 orders or more for
REPORT # 2. If you do not, continue advertising or
sending e-mails until you do.

** Once you have received 100 or more orders for
Report # 2, YOU CAN RELAX, because the system is
already working for you, and the cash will continue
to roll in ! THIS IS IMPORTANT TO REMEMBER: Every time
your name is moved down on the list, you are placed
in front of a different report.

You can KEEP TRACK of your PROGRESS by watching which
report people are ordering from you. IF YOU WANT TO
GENERATE MORE INCOME SEND ANOTHER BATCH OF E-MAILS
AND START THE WHOLE PROCESS AGAIN. There is NO LIMIT
to the income you can generate from this business !!!
=================================================
FOLLOWING IS A NOTE FROM THE ORIGINATOR OF THIS PROGRAM:
You have just received information that can give you
financial freedom for the rest of your life, with
NO RISK and JUST A LITTLE BIT OF EFFORT. You can
make more money in the next few weeks and months
than you have ever imagined. Follow the program
EXACTLY AS INSTRUCTED. Do Not change it in any way.
It works exceedingly well as it is now.

Remember to e-mail a copy of this exciting report
after you have put your name and address in
Report #1 and moved others to #2 .....# 5 as
instructed above. One of the people you send this
to may send out 100,000 or more e-mails and your
name will be on every one of them.

Remember though, the more you send out the more
potential customers you will reach. So my friend,
I have given you the ideas, information, materials
and opportunity to become financially independent.

IT IS UP TO YOU NOW !

=============MORE TESTIMONIALS===============
My name is Mitchell. My wife, Jody and I live in
Chicago. I am an accountant with a major U.S.
Corporation and I make pretty good money. When I
received this program I grumbled to Jody about
receiving 'junk mail'. I made fun of the whole
thing, spouting my knowledge of the population
and percentages involved. I 'knew' it wouldn't work.
Jody totally ignored my supposed intelligence and
few days later she jumped in with both feet. I made
merciless fun of her, and was ready to lay the old
'I told you so' on her when the thing didn't work.

Well, the laugh was on me! Within 3 weeks she had
received 50 responses. Within the next 45 days she
had received total $ 147,200.00 ......... all cash!
I was shocked. I have joined Jody in her 'hobby'.

Mitchell Wolf M.D., Chicago, Illinois
================================================
Not being the gambling type, it took me several
weeks to make up my mind to participate in this plan.
But conservative as I am, I decided that the initial
investment was so little that there was just no way
that I wouldn't get enough orders to at least get
my money back. I was surprised when I found my
medium size post office box crammed with orders.
I made $319,210.00 in the first 12 weeks.

The nice thing about this deal is that it does not
matter where people live. There simply isn't a
better investment with a faster return and so big.

Dan Sondstrom, Alberta, Canada
=================================================
I had received this program before. I deleted it,
but later I wondered if I should have given it a try.
Of course, I had no idea who to contact to get
another copy, so I had to wait until I was e-mailed
again by someone else......... 11 months passed then
it luckily came again...... I did not delete this one!
I made more than $490,000 on my first try and all the
money came within 22 weeks.

Susan De Suza, New York, N.Y.
=================================================
It really is a great opportunity to make relatively
easy money with little cost to you. I followed the
simple instructions carefully and within 10 days the
money started to come in. My first month I made $20,
in the 2nd month I made $560.00 and by the end of the
third month my total cash count was $362,840.00.
Life is beautiful, Thanx to internet.

Fred Dellaca, Westport, New Zealand
=================================================

ORDER YOUR REPORTS TODAY AND GET STARTED ON YOUR
ROAD TO FINANCIAL FREEDOM !

=================================================
If you have any questions of the legality of this
program, contact the Office of Associate Director
for Marketing Practices, Federal Trade Commission,
Bureau of Consumer Protection, Washington, D.C.

This message is sent in compliance of the proposed
bill SECTION 301, paragraph (a)(2)(C) of S. 1618.

* This message is not intended for residents in the
State of Washington, Virginia or California,
screening of addresses has been done to the best
of our technical ability.

6997UQjg3-669fCJn5819fHns7-296kNJT5019vwoU0-971KltO6313sGjr5-646xnHV8706VlTT1-476vl77


From daemon Sun Oct 6 19:58:22 2002



From: arrobert-at-mtu.edu (by way of Ask-A-Microscopist)
Date: Sun, 6 Oct 2002 19:53:39 -0500
Subject: Ask-A-Microscopist:TEM Specimen Prep Help...Staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (arrobert-at-mtu.edu) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Sunday,
October 6, 2002 at 19:39:15
---------------------------------------------------------------------------

Email: arrobert-at-mtu.edu
Name: Amy Roberts

Organization: Michigan Technological University

Education: Undergraduate College

Location: Houghton, Mi United States of America

Question: I am in search of specimen preparation procedures for
transmission electron microscopy for a gram negative and a gram
positive bacteria. Could anyone point me to a journal or publication
with this information?

---------------------------------------------------------------------------


From daemon Mon Oct 7 07:19:02 2002



From: Beauregard, Paul A. :      pabeauregard-at-ppg.com
Date: Mon, 7 Oct 2002 08:08:28 -0400
Subject: Sorvell MT2-B Ultra microtome manual request.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I would like to ask if anyone could fax me a copy of the operating instructions for a
Sorvell MT2-B Ultra microtome.
It is motorized if that matters.

This is for a lady that is not on the list and that works at another research center.

Thank you in advance for your help and time.

724-325-5105 My FAX

Sincerely,

Paul Beauregard
Senior Research Associate
PPG Industries
Monroeville Technical Center
Electron Microscopy
440 College Park Drive
Monroeville, PA 15146

} To calculate the size of one micron in ANY microscopic photograph, one only has to know the magnification. Take the mag and convert it to KX. That number is the number of millimeters that represents one micrometer on your photo. For example, 81385X = 81.385KX and 1micron is 81.385 mm. 516X = 0.516KX and one micron is .516 mm. 1,300,000X = 1300KX and 1300 mm is 1 Micron or 130MMs is 0.1 microns or 13 mm is 10NMs or 100Å.
}
}
}


From daemon Mon Oct 7 08:12:12 2002



From: michael shaffer :      rarewolf-at-roadrunner.nf.net
Date: Mon, 7 Oct 2002 10:33:52 -0230
Subject: RE: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Scott writes ...

} I have been coating samples via electron beam evaporation to a
} thickness of
} 20-25 nm as determined by blue interference colors on polished brass.
} Recently the mineral guys lost their coater so I have also been coating
} their material to the same blue color they said they used as
} well. They used
} resistance heating.
}
} In examining those samples they noticed their EDS/WDS results appearing a
} little low so we went ahead and coated the standards as well.
}
} More digging has revealed they coat with graphite rods whereas I have been
} using amorphous C rods. They swear the coats are thicker and looking at
} theirs-vs-mine it appears so BUT we both evaporated to the same blue color
} on brass.
}
} Could there be a difference in the optical properties of the graphite and
} amorphous rods??

I can't imagine any difference once the carbon has been evaporated ...
i.e., there may be a difference with respect to resistance and temperature
of evaporation (sublimation?) between vitreous carbon and graphite, but once
the transformation happens, carbon is carbon.

I happen to use the red-to-blue transistion, because it is very
distinguishable color change and should result in a coating thickness of
22nm +- 1nm.

cheerios ... shAf :o)
Avalon Peninsula, Newfoundland
www.micro-investigations.com (in progress)




From daemon Mon Oct 7 08:29:29 2002



From: Rik Brydson :      mtlrmdb-at-leeds.ac.uk (by way of MicroscopyListserver)
Date: Mon, 7 Oct 2002 08:23:19 -0500
Subject: Electron Microscopy and Analysis Conference 2003

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


***************************************************************
First Announcement

Electron Microscopy and Analysis Conference 2003
The University of Oxford
3 - 5 September 2003

Organised by the Electron Microscopy and Analysis Group of the Institute of
Physics
Co-sponsored by the Royal Microscopical Society

Scientific Content
Electron Microscopy is going through the most revolutionary period since its
invention, with leaps in performance of sources, imaging lenses and
spectrometers all at roughly the same time. Scanning probe techniques are
also maturing into useful analytical tools. This is an ideal time for the
microscopy community to appraise when, how and why we use different
techniques to gain different scientific insights. The conference will
address all aspects of electron and scanning probe microscopy and
associated spectroscopy including:

o New Instrumentation - including Aberration Correction and
Monochromation
o New Information from New Instrumentation - Applications
Highlighting Novel Techniques
o Advances in Nanoanalysis - EDX, PEELS and Other Analytical
Techniques
o State-of-the-art SEM, Scanning Probe Microscopy and
Surface Science
o High-resolution Electron Microscopy and Electron
Crystallography
o Advances in Sample Preparation

Applications of microscopy and related techniques will be divided into
sessions according to demand.
These will include experiments in:

o Biogenic Materials
o Catalysis
o Ceramics and Electroceramics
o Composites
o Environmental Samples
o Ferrous and Non-Ferrous Metals
o Fullerenes, Hard Carbon and Related Materials
o Intermetallics
o Interfaces and Surfaces
o Magnetic Materials
o Sensor Materials
o Semiconductors
oSuperconductors

Sessions of oral and poster presentations will be arranged, and, where
possible, associated with invited
keynote speakers. Papers from post-graduate students are particularly
encouraged; student bursaries
will be available through the Institute of Physics EMAG Group.

Advanced School
The Advanced School provides an opportunity for postgraduate students,
research assistants and others with an interest to learn about a range of
advanced imaging techniques in electron and related microscopies techniques
from experts in the field. The school will involve a combination of tutorial
lectures and hands-on demonstrations in the Department of Materials, using
the wide range of sophisticated instrumentation available there. The
lecturers/demonstrators will be leading exponents in the techniques
concerned. There will be considerable opportunity for interaction. To
maximise the effectiveness of the School, numbers will be limited to about
20.

Trade Exhibition
Accompanying the conference there will be a trade exhibition where delegates
and visitors will be able to see and discuss the latest products and
developments.

Social Programme
The social programme will include a welcome reception, an Exhibitor Buffet
and a conference dinner. Further details will be available in due course.

Conference Venue
The scientific sessions, the Exhibition and the poster presentations will
take place at the Oxford University Examination Schools. For more
information about how to get to Oxford please visit the Oxford
University web site at http://www.ox.ac.uk/aboutoxford/how.shtml

Important Dates
January 2003 all for Papers
21 March 2003 Deadline for Abstract Submission
May 2003 Registration Mailing
18 July 2003 Deadline for Registration


Organising Committee

EMAG Chairman R M D Brydson
Local Committee Chairman A Petford-Long/J Hutchison
Programme Organiser A Bleloch
Proceedings Editor S McVitie
Exhibition Organiser R Doole
Advanced School Organiser D J Cockayne

General Enquiries
Jasmina Bolfek-Radovani, The Institute of Physics, 76 Portland Place, London
W1B 1NT
Tel +44 (0)20 7470 4800 Fax: +44 (0)20 7470 4900
Email: conferences-at-iop.org
http://physics.iop.org/IOP/Confs/EMG/

Exhibition Enquiries
Ron Doole, EMAG03, Department of Materials, University of Oxford, Parks
Road, Oxford., OX1 3PH
Tel: + 44 (0)1865 273701 Fax: + 44 (0)1865 283333
Email:ron.doole-at-materials.oxford.ac.uk


Media Registration
Accredited journalists are welcome to attend all or part of this meeting
free-of-charge. To qualify for free
registration please contact Alice Bows, Press Officer, Public Affairs
Department, Tel: +44 (0)20 7470 4800, Fax: +44 (0)20 7470 4848, Email:
alice.bows-at-iop.org.



From daemon Mon Oct 7 09:04:55 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Mon, 7 Oct 2002 10:03:28 -0400
Subject: Re: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Sergey & Co.;

In an analysis I did several years ago I wanted to place a controlled amount
of carbon on a surface and measure the ultimate strength of a gold to gold
weld in a microcircuit. We had great difficulty with this since we had no
method to know, with any accuracy, the real deposition thickness. Our
target was 100 angstroms, with increments up to 500. We attempted to use
Auger, however, with a carbon species and the fact that the first 50
angstroms in the Auger survey are thrown out, we never actually determined
what the final result was. Would AFM have been a better choice for this
measurement?

Peter Tomic
Anadigics, Inc.

-----Original Message-----
} From: Sergey Ryazantsev [mailto:sryazant-at-ucla.edu]
Sent: Friday, October 04, 2002 9:47 PM
To: Microscopy-at-sparc5.microscopy.com


Tobias, good question

In specification it indicates 0.01 nm accuracy as far as I remember, but as
I mentioned in my original posting, it seems to me that I do have
approximately 10% difference in the film thickness even with that smart
thickness monitor. In general, I am happy with that instrument. It's very
solid and sensitive enough in 10-20 A scale. Sergey

At 04:15 PM 10/4/02, you wrote:
} Dear Sergey,
} Does your thickness monitor really measure down to 0.01 nm
} accuracy?? That's amazing.
}
} Tobias
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Mon Oct 7 10:06:56 2002



From: Anthony J. Garratt-Reed :      tonygr-at-mit.edu
Date: Mon, 07 Oct 2002 10:49:57 -0400
Subject: NESM Fall Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html




To all:

The Fall meeting of NESM (New England Society for Microscopy) will be held
on Tuesday, October 15th at M.I.T. in Cambridge, MA.

The program will be devoted to Raman Spectroscopy, and the schedule is as
follows:

5:30 - 6:00pm Registration
6:00 - 7:15pm Buffet Supper
7:15 - 8:45pm Technical sessions

7:15pm - Raman Microspectroscopy and Imaging
David S. Himmelsbach, United States Department of Agriculture Agricultural
Research Service
Richard B. Russell Research Center, Athens, Georgia, 30677

8:00pm - Raman Spectroscopy on Isolated Carbon Nanotubes
Shin Grace Chou, Center for Materials Science and Engineering,
Massachusetts Institute of Technology, Cambridge, MA 02139


Place: Stratton Student Center (Building W20), Twenty Chimneys Lounge
(3rd. floor), 84 Massachusetts Avenue, Cambridge.

Travel: A map and directions are on the Web at
http://whereis.mit.edu/bin/map?state=0&pri.x=281&pri.y=123
Parking will be extremely difficult because of on-going road construction
on Vassar Street (the main location of on-street parking). If you park in
any MIT lot, even in the evening, you will be ticketed unless you have an
MIT sticker. There is a public lot on the corner of Vassar St. and Mass.
Ave. that may be available. There is a public garage behind the Legal
Seafoods Restaurant in Kendall Square (10 mins. walk). There is another
garage several blocks North just off Mass. Ave. (2 blocks beyond the NECCO
building). It should also be possible to find on-street parking, but be
prepared to walk.

If possible, we recommend you use public transport. The number 1 bus from
Harvard to Dudley stops on Mass Ave outside the student center. The Red
Line stops at Central and Kendall are each 10 minutes walk (use Central if
you come from Harvard or Alewife, Kendall if you come from Boston or points
South.)

Registration for members is $5.00 and for non-members is $20.00 (this
includes a one-year membership in NESM). The deadline for registration is
Friday, October 11th. Please contact Paul Bain, NESM Treasurer,
(Paul_Bain-at-hms.harvard.edu) if you are planning to attend.



** Anthony J. Garratt-Reed
** MIT Room # 13-1027
** 77 Massachusetts Avenue
** Cambridge, MA 02139-4307
** USA
**
** Phone: (+) 1-617-253-4622
** Fax: (+) 1-617-258-6478
**




From daemon Mon Oct 7 11:20:04 2002



From: Jinguo Wang :      jqw11-at-psu.edu
Date: Mon, 07 Oct 2002 12:10:08 -0400
Subject: LN2 and/or heating holder for a 2000FX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi, Fellows,

We are interested in buying a used 2000FX LN2 and/or heating holder, If
anyone would like to sell those used holders please let me know.

Thanks,

Jinguo Wang, Ph.D
The Pennsylvania State University
Materials Research Institute
194 Materials Research Institute Building
University Park, PA 16802
Tel: (814) 865-9285
Fax: (814) 863-8561, (814) 863-0637
email: jqw11-at-psu.edu



From daemon Mon Oct 7 12:56:36 2002



From: JHoffpa464-at-aol.com
Date: Mon, 07 Oct 2002 13:49:15 -0400
Subject: Re: Sorvell MT2-B Ultra microtome manual request.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


mt-2s are so easy, why would she need a manual?


From daemon Mon Oct 7 13:23:04 2002



From: Marcia A. Miller, Ph.D. ABMM :      Marcia-at-dbts.uicomp.uic.edu
Date: Mon, 7 Oct 2002 13:21:23 -0500
Subject: TEM: Fusarium & Penicillium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


We use a great deal of carbon and graphite daily for making our substrates.
Some in our lab prefer graphite, some carbon.
If you examine the crystalline structure you'll find carbon is more random
(less crystalline). Graphite requires more current in the Ladd evaporators.
Carbon is more expensive.
We offer carbon and high quality graphite. Based on our experience and
customer feedback we are of the opinion that excellent results can be
attained with carbon or high quality graphite.

John Arnott

Ladd Research
83 Holly Court
Williston, VT 05495

On-line Catalog: http://www.laddresearch.com

tel: 1-802-658-4961(anywhere) or 1-800-451-3406(US)
fax: 1-802-660-8859
e-mail: sales-at-laddresearch.com
----- Original Message -----
} From: "Ritchie Sims" {r.sims-at-auckland.ac.nz}
To: "Sergey Ryazantsev" {sryazant-at-ucla.edu} ;
{Microscopy-at-sparc5.microscopy.com}
Sent: Sunday, October 06, 2002 7:12 PM



Ladies/Gentlemen:

I am searching for TEMs of two fungi; one is our newly
discovered Penicillium that is related to several listed below.
Need TEM's with good ultrastructure and labeling of
internal structures. Will appreciate help by listing references,
text books, personal electronmicrographs, etc. Need information
for publication purposes.

1) Fusarium verticilloides (moniliforme) or related species.

2) Penicillium species related to new species identified:

a. Penicillium restrictum
b. Penicillium dimorphosporum
c. Penicillium striatisporum

If TEM's of above Penicillium species not available,
then electronmicrographs of monoverticillate, nonvesiculate
penicilli with short stipes.

Thank you.

e-mail: mam-at-uic.edu






From daemon Mon Oct 7 13:29:04 2002



From: Nancy.P.Piatczyc-at-williams.edu
Date: Mon, 07 Oct 2002 14:23:53 -0400
Subject: mass spec upgrade

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear list,

This is off-topic, many apologies, but I am stumped. I am not familiar
with mass spectrometers, but have been asked to get information on
upgrading a 1070's vintage mass spec - quadrupole. The manufacturer is
Nuclide. They want to send the detector output to a PC. Many thanks
for any info from vendors or individuals. Nancy

Nancy Piatczyc
Bronfman Science Center
Williams College
Williamstown, MA 01267






From daemon Mon Oct 7 14:03:36 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Mon, 07 Oct 2002 11:54:19 -0700
Subject: RE: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Peter

Actually, only one direct way to measure REAL thickness I do know: it's
embedding in the plastic and cross-sectioning strongly PERPENDICULAR. All
other ways are arbitrary. As I mentioned in original report, I do have 10%
difference from experiment to experiment for carbon. Heavier elements may
be measured, perhaps, more precisely... Sergey

At 07:03 AM 10/7/02, you wrote:
} Sergey & Co.;
}
} In an analysis I did several years ago I wanted to place a controlled amount
} of carbon on a surface and measure the ultimate strength of a gold to gold
} weld in a microcircuit. We had great difficulty with this since we had no
} method to know, with any accuracy, the real deposition thickness. Our
} target was 100 angstroms, with increments up to 500. We attempted to use
} Auger, however, with a carbon species and the fact that the first 50
} angstroms in the Auger survey are thrown out, we never actually determined
} what the final result was. Would AFM have been a better choice for this
} measurement?
}
} Peter Tomic
} Anadigics, Inc.
}
} -----Original Message-----
} From: Sergey Ryazantsev [mailto:sryazant-at-ucla.edu]
} Sent: Friday, October 04, 2002 9:47 PM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: Re: SEM Carbon Coating thickness
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Mon Oct 7 14:37:23 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Mon, 7 Oct 2002 15:36:14 -0400
Subject: RE: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Sergey;

Even with a great FESEM as we have, I would be hard pressed to image 100
angstroms of anything with a 10 angstrom accuracy. The FESEM sometimes can
do 15 angstroms on an ideal sample on a good day, but I could never resolve
10A/100A. By "perpendicular" x-sectioning, are you referring to mount lap
and polish, microtome? Also, what is a "Smart Thickness Monitor" you mention
below? Maybe I need one. In short, how does it do the measurement, a stylus
like AFM, optical method?

Regards,

Peter

Ps. I would imagine the Los Angeles smog would put 100 angstroms of soot
down on a glass slide in short order, unlike pristine New Jersey.

-----Original Message-----
} From: Sergey Ryazantsev [mailto:sryazant-at-ucla.edu]
Sent: Monday, October 07, 2002 2:54 PM
To: Peter Tomic; Microscopy-at-sparc5.microscopy.com

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu




From daemon Mon Oct 7 14:53:44 2002



From: Scott Whittaker :      Whittaker.scott-at-nmnh.si.edu
Date: Mon, 07 Oct 2002 15:42:03 -0400
Subject: Carbon vs graphite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Ok here's a new twist. It was kinda slow Friday and we ended up running
several different rods and simulated samples again coated to blue on brass.
My brass washer was the same color as theirs. The consensous upstairs is
still that that my coatings are substantially thicker and looking at the
samples I agree they are darker.

They evaporated carbon and sometimes graphite via resistance. I use e-beam
evaporation, and graphite rods take about 5 minutes, heating the head to
cherry in color so is not really an option. Could the different thickness be
due to different density of the films? The e-beam evaporated material in
theory has a finer more uniform structure with less space, thus the film is
more "dense" There would be more carbon in the film per unit thickness so
when I coat to the blue 25nm, more carbon is present.

Resistance evaporation is less controlled and there is less carbon in the
film structure, but still colors brass blue at 25nm.

The other option is that a denser film optically reflects differently and
the blue is something like 40nm or so. I had ruled that out since I have a
thickness monitor and it is reading correctly.

Anyway, just some more thoughts to toss around. I never really paid any
attention to the carbon films and have already learned much from the
discussion so far. Thanks


From daemon Mon Oct 7 15:01:55 2002



From: Nancy.P.Piatczyc-at-williams.edu
Date: Mon, 07 Oct 2002 15:55:39 -0400
Subject: mass spec correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Opppps, typo, it's a 1970's vintage that needs an upgrade. Nancy




From daemon Mon Oct 7 15:31:33 2002



From: Philip Oshel :      peoshel-at-facstaff.wisc.edu
Date: Mon, 07 Oct 2002 15:21:48 -0500
Subject: RE: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


What is the actual molecular structure of an "amorphous" carbon rod?
Is it really amorphous? Microcrystalline? Something else?
Given the crystal form of graphite, I suspect that it is a much more
efficient generator of bucky balls, nanotubes, and suchlike
structures. This could explain the problems with film thickness and
consistency from graphite evaporation vs (amorphous) carbon rods. How
would the bucky balls, nanotubes, etc. affect the electrical
properties of a carbon film, and thus its behavior in an electron
beam?
Or is this all wrong and irrelevant?

Has anyone done any AFM/STM on carbon films deposited by these
different methods?

Phil

} Scott writes ...
}
} } I have been coating samples via electron beam evaporation to a
} } thickness of
} } 20-25 nm as determined by blue interference colors on polished brass.
} } Recently the mineral guys lost their coater so I have also been coating
} } their material to the same blue color they said they used as
} } well. They used
} } resistance heating.
} }
} } In examining those samples they noticed their EDS/WDS results appearing a
} } little low so we went ahead and coated the standards as well.
} }
} } More digging has revealed they coat with graphite rods whereas I have been
} } using amorphous C rods. They swear the coats are thicker and looking at
} } theirs-vs-mine it appears so BUT we both evaporated to the same blue color
} } on brass.
} }
} } Could there be a difference in the optical properties of the graphite and
} } amorphous rods??
}
} I can't imagine any difference once the carbon has been evaporated ...
} i.e., there may be a difference with respect to resistance and temperature
} of evaporation (sublimation?) between vitreous carbon and graphite, but once
} the transformation happens, carbon is carbon.
}
} I happen to use the red-to-blue transistion, because it is very
} distinguishable color change and should result in a coating thickness of
} 22nm +- 1nm.
}
} cheerios ... shAf :o)
} Avalon Peninsula, Newfoundland
} www.micro-investigations.com (in progress)

--
Philip Oshel
Supervisor, BBPIC microscopy facility
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706 - 1284
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)


From daemon Mon Oct 7 18:41:23 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Mon, 07 Oct 2002 16:30:37 -0700
Subject: RE: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


When carbon (or any other material) evaporated (sublimated for carbon?) as
I understand, it's not mono-atomic "vapor". It contains atomic clusters of
different size. Those clusters may be produced during "evaporation" or
later in the "vapor" clouds. I could imagine that evaporation conditions,
the density of cloud etc may shift local conditions in the favor of some
carbon state. Condensation conditions on the surface may also vary
(temperature etc). So, to me it's easy to imagine how it happens that
carbon film is heterogeneous (mica steps could initiate some specific
condensation). I was thinking to do AFM on different "carbons". It should
be done in the vacuum environment I guess. Unfortunately, I don't have
access to such instrumentation. Sergey

At 01:21 PM 10/7/02, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From daemon Tue Oct 8 07:15:01 2002



From: Tony Bruton :      Bruton-at-nu.ac.za
Date: Tue, 08 Oct 2002 14:04:48 +0200
Subject: Oil mist problem with JEOL 100CX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Greetings All

I have been told that the problem we are currently experiencing with
our ~ 1980 JEOL 100CX has been seen before.

We would very much appreciate any information or advice.

We are getting an accumulation of oil mist (DP oil ?) on the inside of
the viewing chamber glass in the specimen chamber of our 100CX. Where is
it coming from, where do I start looking for the fault ? We no sooner
clean it off than it is back again.

Look forward to hearing from listers.

Tony

Tony Bruton
Head, Centre for Electron Microscopy
University of Natal, Pietermaritzburg
Tel +27 (0) 33 260 5155
Fax +27 (0) 33 260 5776
Mob. 082 782 9640
website: www.nu.ac.za/microscopy.asp
Email: bruton-at-nu.ac.za
Postal address;
Private Bag X01,
Scottsville, 3209
KwaZulu-Natal
South Africa


From daemon Tue Oct 8 07:46:54 2002



From: Robert Wieland :      wieland-at-me.udel.edu
Date: Tue, 8 Oct 2002 08:38:57 -0400 (EDT)
Subject: Re: SEM Carbon Coating thickness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Excuse me if this is common knowledge...
Thirty years ago, the organization I worked for was fabricating carbon
"bottles" to contain cadmium sulphide to be heated for vacuum deposition.
We ran into the problem that commercial "graphite" contains several
percent boron, put in to make the stuff machinable. Since we didn't want
boron involved in our process, we had to specify "nuclear-grade" graphite,
intended for reactor parts where they didn't want any boron, either.
This turned out to be hard enough to get that we laid in a ten-year supply
when we finally found some.
Are "graphite" evaporator rods boron-free? If not, could that be
responsible for the differences reported? I assume amorphous carbon is
pure.



Robert Wieland wieland-at-me.udel.edu
The very concept of human governance is a moral dilemma:
If the people are good, it is a mistake to create authorities over them;
If they are not good, it is a mistake to create authorities out of them.

You can easily peg people by listening to them recount a hand:
winners talk about how they played the cards they were dealt;
non-winners, about how they should have played the cards they were dealt;
and losers, about the cards they should have been dealt.

R F Wieland wieland-at-me.udel.edu 39.68N 75.74W
Icom R75 Heathkit GR-81 Inverted-L in the attic



From daemon Tue Oct 8 09:12:11 2002



From: Kathryn Zeisler-Mashl :      kzeisler-at-wpi.edu
Date: Tue, 8 Oct 2002 10:11:28 -0400
Subject: TEM - 100C double tilt holder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,
We are looking for a double tilt (tilt-tilt) holder for a Jeol 100C/100CX to
examine metallurgical specimens. If you have one available, please reply to
me directly (kzeisler-at-wpi.edu).

Thank you very much.
Kate Zeisler-Mashl

-----------------------------------------
Kathryn L. Zeisler-Mashl, Ph.D.
Research Assistant Professor
Materials Science and Engineering Program
Worcester Polytechnic Institute
100 Institute Road, Worcester, MA 01609
T: (508)831-6025 F: (508)831-5178
kzeisler-at-wpi.edu





From daemon Tue Oct 8 09:16:41 2002



From: Dorota Wadowska :      wadowska-at-upei.ca
Date: Tue, 08 Oct 2002 11:09:43 -0300 (ADT)
Subject: EM job classificcation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Listservers!
I have a few questions that I would like to ask you If you feel you
know the answer please reply. My university together with local
union reclassified staff positions according to Aiken plan. Needless
to say my position was downgraded.
My questions are:
1. Can EM supervisory position be compared to any on campus
staff positions (university has Department of Sciences and
Veterinary Medecine)? I supervise small multi--user lab, we have 2
TEM, and we do research, diagnostics, teaching (under- and
graduate level) in the field of biological sciences. No fancy staff like
X-ray analysis or cryo.
2. What in your opinion is a necessary minimum experience
required to perform this job ? I am talking about accepting the
position and immediately getting into the stream of work.
3. What in your opinion is the minimum level of education that is
required to perform this job?

You can reply directly to me. Thanks in advance
Frustrated
Dorota


From daemon Tue Oct 8 10:19:21 2002



From: Drew Price :      dprice2-at-sigecom.net
Date: Tue, 8 Oct 2002 11:00:43 -0500
Subject: McCrone Research Institute

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Boron is in the manufacturing process. It's probably in the PPB range in
the graphite rods. Some graphite rods may be boron free but you'd have to
assume there is some boron in most graphite rods.

Disclaimer: Ladd sells a wide variety of electron microscope related
products.

John Arnott

Ladd Research
83 Holly Court
Williston, VT 05495

On-line Catalog: http://www.laddresearch.com

tel: 1-802-658-4961(anywhere) or 1-800-451-3406(US)
fax: 1-802-660-8859
e-mail: sales-at-laddresearch.com
----- Original Message -----
} From: "Robert Wieland" {wieland-at-ME.UDel.Edu}
To: "John Arnott" {ladres-at-worldnet.att.net}
Cc: "Microscopy Listserver" {Microscopy-at-sparc5.microscopy.com}
Sent: Tuesday, October 08, 2002 8:38 AM


I am scheduled to attend McCrone's applied polarized light microscopy course
later this month. I have a medical background consisting of gross human
anatomy and physiology, a limited science background, law enforcement and
legal experience, but I currently do not work in a science field. A friend
of mine recently provided me with some "hands-on" with a polarized light
microscope.

How else can I prepare for this course of instruction? I am paying for this
training myself, and I want to get the most out of it.

Any input or advice would be greatly appreciated.

Thank you,

Drew Price
dprice2-at-sigecom.net




From daemon Tue Oct 8 12:38:05 2002



From: Barbara Maloney :      maloneyb-at-fiu.edu
Date: Tue, 08 Oct 2002 13:44:37 -0400
Subject: turbo pump carbon coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

More on the discussion on carbon/graphite rods used in EM labs for
evaporation:

To understand what the situation really is, you have to go back to the
"starting" material, which differs between the different manufacturers of
the final product. The starting material could be either coke, coal, or
what is called "pitch". Since this can come from different geographical
locations, the level of impurities varies, including the level of boron
which is everywhere.

The first step is the converstion of the pitch to carbon and then the second
step is the conversion of the carbon to graphite. At least that is the
process used in the production of the SPI Supplies offered rods. The
problem arises because boron very easily forms a carbide and that carbide is
very difficult to purify out of the final product.

In years gone by, and before analytical instrumentation had the low
detection limits of today's instrumentation, it was indeed common to talk
about carbon or graphite as being "boron free". Of course, nothing is
ever zero, we can only talk about detection limits and say that if present,
it is less than some amount.

The instrumentation used for the control of boron in the SPI Supplies carbon
and graphite rod products has a detection limit of slightly more than 1 ppm
and we keep our product to a boron level of less than 2 ppm.

The purification cost is a major cost element of the delivery of high purity
carbon or graphite rods; not all applications require this kind of purity.
The way products are presented, either in catalogs or websites, might
suggest they are all the same but they are not all the same. I can not
address the purity of the carbon or graphite rods sold by others, only that
offered by SPI Supplies. But certainly this level of purity of boron is
well beyond that needed by anyone using these products in an EM lab
application.

} From our experience, the biggest reason why different rods from different
sources evaporate differently is due to differences in the density of the
grahpite rods.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================





------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America To
Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com On-Line
Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
-----------------------------------------------------------------------.


Boron is in the manufacturing process. It's probably in the PPB range in
the graphite rods. Some graphite rods may be boron free but you'd have to
assume there is some boron in most graphite rods.

Disclaimer: Ladd sells a wide variety of electron microscope related
products.

John Arnott

Ladd Research
83 Holly Court
Williston, VT 05495

On-line Catalog: http://www.laddresearch.com

tel: 1-802-658-4961(anywhere) or 1-800-451-3406(US)
fax: 1-802-660-8859
e-mail: sales-at-laddresearch.com
----- Original Message -----
} From: "Robert Wieland" {wieland-at-ME.UDel.Edu}
To: "John Arnott" {ladres-at-worldnet.att.net}
Cc: "Microscopy Listserver" {Microscopy-at-sparc5.microscopy.com}
Sent: Tuesday, October 08, 2002 8:38 AM


Hi group - anybody out there that currently uses a carbon coater with a
turbo pump? Looks like our Pelco carbon coater died and parts are hard
to find.
Can any of you recommend a carbon coater that you are happy with?
Thanks
Barb



From daemon Tue Oct 8 12:54:21 2002



From: Barbara Maloney :      maloneyb-at-fiu.edu
Date: Tue, 08 Oct 2002 13:45:49 -0400
Subject: cold stage and SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi didn't really get a response from my last query - anybody currently
using a CL with SEM? We are looking to upgrade our system and get a cold
stage.
Thanks
Barb



From daemon Tue Oct 8 13:28:19 2002



From: Geoff McAuliffe :      mcauliff-at-umdnj.edu
Date: Tue, 08 Oct 2002 17:20:24 -0400
Subject: Re: McCrone Research Institute

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Tony,

This problem might be caused by several conditions. First, determine what
oil is
present in the projection chamber. Possibilities are: Santovac (DP),
Hydrocarbon (PVP), and air compressor oil, which could be either Hydrocarbon
or an Ester, depending on compressor type. The latter might be irrelevant in
case that you have replaced original JEOL compressor with a dry compressor,
however: an old compressor oil could be present in the pneumatic
lines/valves for years, and might eventually show up on the vacuum side of
pneumatic valves upon moveable seal(s) failure.

I use a highly volatile solvent (such as Pentane), others may work too.
Pentane dissolves Hydrocarbon oil very well and fast, Santovac very poorly,
and Ester oil to some degree. In addition, compressor oil (especially one
sitting in the lines and valves for years, and especially Ester oil) has
deep red/brown color, clearly visible even in microscopic droplets.

Remove a small polished metal part with oil mist visible on it from the
contaminated area. Observe oil mist under stereo microscope, see the color,
try consistency of the droplets with the needle. Santovac is much more
viscous, especially in a cool room. Rinse that part briefly in
Pentane, see if some droplets have disappeared, or were reduced in size more
than others.

Many scenarios are possible, including following:
a) ODP(s) cooling failed (clogged water lines) - Santovac in the
column/viewing chamber. Likely scenario.

b) Pneumatically actuated vacuum valves acting slowly (damage, dirt,
deformed moving seals, compressor oil and water in pneumatic cylinders)- PVP
Hydrocarbon oil or/and Santovac are blown into vacuum when valves switching
for venting or rough pumping is delayed. Likely scenario.

c) Damaged seals in the pneumatically actuated vacuum valves- compressor oil
in the column/viewing chamber. BTW, do you drain oil and water from air
compressor tank regularly? Do you
have water mist filter installed between compressor and a TEM? Likely
scenario.

d) Damaged PVP- Hydrocarbon oil makes it's way into the ODP, and from there
into column/viewing chamber. Unlikely scenario.

e) Pirani gauges or circuits are defective- vacuum sequences are out of
order. Connect stand-alone vacuum gauge to a TEM (multiple ports) and verify
switching points/thresholds. Unlikely scenario.

Unfortunately, an easy fix is unlikely for a problem like this. I recommend
spending some time and conduct an investigation, with vacuum system diagram
in mind. This will help to trace problem more accurately, and avoid
unnecessary disassembly.

Possible remedies may include cleaning/re-packing pneumatically actuated
vacuum valves, cleaning pneumatic lines, cleaning an ODP and replacing the
Santovac, replacing worn out air compressor, replacing/servicing PVP.

Vitaly Feingold
Scientific Instruments and Applications
2773 Heath Lane, Duluth GA 30096
(770)232-7785 ph.
(770)232-1791 fax
(678)467-0012 mobile

This message is made of 100% recycled electrons.

This address can not receive messages larger than 15 kb without prior
notification.

----- Original Message -----
} From: Tony Bruton {Bruton-at-nu.ac.za}
To: {Microscopy-at-sparc5.microscopy.com}
Sent: Tuesday, October 08, 2002 8:04 AM


There is an excellent book on the subject by F. Donald Bloss called "Optical
Crystallography" and it is available from the Minerological Society of
america for about $25.00.

Don
----- Original Message -----
} From: "Drew Price" {dprice2-at-sigecom.net}
To: {Microscopy-at-sparc5.microscopy.com}
Sent: Tuesday, October 08, 2002 12:00 PM


Why not ask the people who run the course?

Geoff

Drew Price wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} I am scheduled to attend McCrone's applied polarized light microscopy course
} later this month. I have a medical background consisting of gross human
} anatomy and physiology, a limited science background, law enforcement and
} legal experience, but I currently do not work in a science field. A friend
} of mine recently provided me with some "hands-on" with a polarized light
} microscope.
}
} How else can I prepare for this course of instruction? I am paying for this
} training myself, and I want to get the most out of it.
}
} Any input or advice would be greatly appreciated.
}
} Thank you,
}
} Drew Price
} dprice2-at-sigecom.net

--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************




From daemon Tue Oct 8 18:33:07 2002



From: HIS HIGHNESS WALID S. AL SAID :      walid_al1-at-mail.com
Date: Wed, 09 Oct 2002 00:23:21
Subject: CLAIMS/INVESTMENTS FROM HIS HIGHNESS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I am HIS HIGHNESS WALID SULMAN AL SAID, Emir of the state of
Bahrain.

I got to know you through a foreign trade office in London
during my official trip last year.

I am making this contact to you to inquire from you if you can
assist me manage some investment for me in your
country.Actually all I want from you is to be my manager
abroad.I would like you to invest in Properties(Real Estate)
and company shares and other lucrative ventures of your
interest for me.

I have Forty Five Million United States Dollars
(US$45,000,000.00)into a VAULT with a Security Finance Company
abroad,but due to many commitments I can not travel out of my
domain for the claim,hence I contacted you to assist me make
the claim on my behalf.

I will provide you with all the necessary documents to enable
you represent me as my business manager to make the claim by
signing the necessary document for onward transfer of the money
into your account which you will use for the investment.

If you agree to assist me by representing me make the claim
with the Security Company, I will now contact the Security
Finance Company to inform them about your person as my
business manager, that they should give you the necessary
attention needed.

For your assistance in this transaction I will offer you 10% of
the total amount, I have also set aside 2% for expenses, while
88% left will be invested into buying of Properties (Real
Estate) company shares and other profit oriented ventures. I
have taken this decision to invest abroad due to restriction
over currency out flow in my country.

Upon your Acceptance I will require the following from you:

1. A letter of Assurance that you will invest and manage wisely
on my behalf.

2. Your telephone and fax numbers to enable the Security
Finance Company communicate with you for the claim. Your
mobile telephone will be of great assistance as well.

For security reasons, I would want every communication between
us in this transaction be confidential. Always reach me via E-
mail for I may not be chanced to speak with you on telephone.

In the later days we will be able to meet for us to discuss
issues of mutual understanding in respect to our investment
project.

Looking forward to a fruitful business relationship with you.

Best regards,

His Highness,
WALID SULMAN AL SAID(Emir of the State of Bahrain).




From daemon Tue Oct 8 20:20:52 2002



From: Damian Neuberger :      neuberger1234-at-attbi.com
Date: Fri, 8 Nov 2002 20:09:10 -0600
Subject: High Speed Video Camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Listers,

I have limited experience with video recording so I may not be giving enough
information; also limited for other reasons. I am in the market for a high
speed video camera and would like some mfg recomendations. We would like to
record a process that takes about 1 second and view an area that is about
1cm square and have a depth of field about 1cm. In other words, the process
takes place within a one cubic cm volume. Also want a camera that can
record without using strobe illumination. Reasonably high resolution (for
HS Video) 800 by 600 would be nice but probably way too expensive. Maybe 2k
frames per second? Would also perfer something that is "somewhat" portable.
Definitely plan to try it out before purchase.

Any recomendations and information in selecting such an instrument would be
very appreciated.

Damian Neuberger
Baxter Healthcare Corp.
damian_neuberger-at-baxter.com



From daemon Tue Oct 8 20:49:30 2002



From: qualityimages :      qualityimages-at-netrax.net
Date: Mon, 07 Oct 2002 18:42:21 -0400
Subject: Re: mass spec upgrade

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html




-------- Original Message --------






From daemon Tue Oct 8 20:49:35 2002



From: qualityimages :      qualityimages-at-netrax.net
Date: Tue, 08 Oct 2002 18:12:30 -0400
Subject: Re: Oil mist problem with JEOL 100CX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html




-------- Original Message --------






From daemon Tue Oct 8 21:15:23 2002



From: Bob Roberts :      bobrobs-at-cox.net
Date: Tue, 8 Oct 2002 19:07:30 -0700
Subject: Oil mist problem with the JEOL JEM-100CX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Tony,

You have a vacuum leak in the camera chamber and the mist you see is
back-streamed diffusion pump oil. Either the high vacuum valve that isolates
the diffusion pump from the camera chamber when venting and exchanging films
is not closing completely (or has a bad "o" ring seal) OR the chamber door
"o" ring is not sealed and you have a steady stream of oil while pumping the
chamber and under high vacuum pumping.

By the way, there is/was a thin coating of carbon on the inside surface of
the leaded viewing glass. I don't know how many times you've cleaned this
off but you may have to recoat it in a high vacuum carbon evaporator.

Bob Roberts
EM Lab Services, Inc.
449 NW 62nd St
Topeka, Kansas 66617-1780
785.246.1232
www.emlabservices.com


Greetings All

I have been told that the problem we are currently experiencing with
our ~ 1980 JEOL 100CX has been seen before.

We would very much appreciate any information or advice.

We are getting an accumulation of oil mist (DP oil ?) on the inside of
the viewing chamber glass in the specimen chamber of our 100CX. Where is
it coming from, where do I start looking for the fault ? We no sooner
clean it off than it is back again.

Look forward to hearing from listers.

Tony





From daemon Tue Oct 8 22:41:53 2002



From: Gary Gaugler :      gary-at-gaugler.com
Date: Tue, 08 Oct 2002 20:34:51 -0700
Subject: Fwd: CLAIMS/INVESTMENTS FROM HIS HIGHNESS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Well, so much for your perfect spam filter setup.

How can his highness make it through this low level
filter scheme??

gg


} From: "HIS HIGHNESS WALID S. AL SAID" {walid_al1-at-mail.com}
} Date: Wed, 09 Oct 2002 00:23:21
} To: MicroscopyFilteredEmail1-at-sparc5.microscopy.com
} Subject: CLAIMS/INVESTMENTS FROM HIS HIGHNESS
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Wed Oct 9 03:04:44 2002



From: Arthur Day :      ard-at-ansto.gov.au
Date: Wed, 9 Oct 2002 17:54:38 +1000
Subject: Re: Oil mist problem with JEOL 100CX]

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


} Tony,
} I can't say that I'm familiar with the 100CX vacuum system, but I
} recently had a similar problem with a couple of JSM-840s.
}


}
} The second was not so bad, but also having to do with the airlock.
} If the airlock is left open (not pumped) most of the time then the
} airlock vent valve (a small solenoid valve) gets quite hot after a
} while and remains hot (sometimes for years on end). This hardens
} the rubber insert in the plunger so that it will no longer seal
} properly. JEOL operates their airlock on a timer rather than from
} Pirrani readings, so when it times out it says you can transfer the
} specimen. This can lead to a considerable gas pulse and some minor
} backstreaming from the DP before the high vacuum valve closes.
}

Greetings colleagues,

We had precisely the same problem with our JSM-6400 some years ago,
that is, the rubber seal on the end of the plunger in the airlock
vent valve went hard after only a couple of years of operation and
wouldn't seal properly anymore because of the heat from the solenoid
valve. After replacing this we decided to avoid a repeat of the
problem by keeping the airlock under vacuum, thus maintaining the
airlock vent valve in the closed state (solenoid not activated), and
therefore cool for years on end and not hot all of the time. This has
had the added advantage of keeping the inside of the airlock a lot
cleaner and "drier" which should further soften the burden on the
diff pumps when opening it to the main chamber...

This has really cleaned things up for us with no more problems for
over 10 years now, and none ever on our ("similar") newer JSM-6300.


Arthur






Arthur Day, Electron Microscope Unit Phone: 61-2-9717-3457
Ansto Materials Division Fax: 61-2-9543-7179
PMB 1, Menai (Sydney), NSW, 2234 Email: ard-at-ansto.gov.au
Australia www: http://www.ansto.gov.au/


From daemon Wed Oct 9 09:14:05 2002



From: Kent Susan-G10900 :      Susan.Kent-at-motorola.com
Date: Wed, 9 Oct 2002 08:57:56 -0500
Subject: McCrone Applied Polarized Light Microscopy Course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Mr. Price,

I attended the PLM class at the McCrone Research Institute in Chicago IL USA
in 1990. I don't know how much, if any, it has changed since then but I
found it to be an extremely informative class. It was a very intensive week
of lectures, hands-on laboratory sessions, and evening homework
assignments. Their teaching staff has a genuine interest in microscopy and
was eager to share their knowledge and enthusiasm with the students. The
fact that you're spending your own money to pay for it, and also that you
have such a diverse educational background tells me that you're very
motivated to study and should therefore have little or no trouble with it.

If you can continue to have access to your friend's polarized light
microscope after completing the class, I strongly recommend doing so.
Polarized light microscopy is not only a science but an art, and like
similar subjects, requires a lot of practice to maintain and improve one's
skills.

Best regards,
Sue Kent
Senior Staff Failure Analysis Engineer
Motorola
4000 Commercial Avenue
Northbrook IL USA 60062
847-480-6834


{ { { {I am scheduled to attend McCrone's applied polarized light microscopy
course
later this month. I have a medical background consisting of gross human
anatomy and physiology, a limited science background, law enforcement and
legal experience, but I currently do not work in a science field. A friend
of mine recently provided me with some "hands-on" with a polarized light
microscope.

How else can I prepare for this course of instruction? I am paying for this
training myself, and I want to get the most out of it.

Any input or advice would be greatly appreciated.

Thank you,

Drew Price
dprice2-at-sigecom.net} } } } }




From daemon Wed Oct 9 09:14:05 2002



From: Hendricks, Gregory :      Gregory.Hendricks-at-umassmed.edu
Date: Wed, 9 Oct 2002 09:55:42 -0400
Subject: RE: Seeking used SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for a reasonably late model (5-10 years old) SEM for sale.
Please respond to this e-mail address. Sincerely

Greg Hendricks
Core EM Facility Manager
University of Massachusetts Medical School
Worcester, MA

Gregory.Hendricks-at-umassmed.edu




From daemon Wed Oct 9 09:42:16 2002



From: Jim Haley :      haley-at-mvia.com
Date: Wed, 09 Oct 2002 10:33:59 -0400
Subject: Re: High Speed Video Camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Damian,

We design custom high speed image recording systems for the biological,
defense, and industrial applications. The software is very easy to
use. We have various configurations depending on requirements for
resolution, frame rate, and cost.

I'd really need some more information from you such as the MINIMUM
acceptable and IDEAL value as well for:
1.) Horizontal and vertical resolution (either in terms of number of
pixels per inch or overall)
2.) Frame rate - is 2000 a minimum?
3.) Record time - is 1 second a maximum for each experiment?
4.) How many record sessions would you like to be able to store before
archiving previous sessions?

Lighting is always a major consideration when doing high speed imaging
as well. Bear in mind that at 2,000 frames per second the camera will
only have 1/66th of the amount of time normally used by standard RS-170
video to allow photons to be collected by the sensor. Instead of the 33
milliseconds exposure time RS-170 video uses, at 2,000 frames per second
we'll only be able to gather light for 0.5 milliseconds. This is one of
the major considerations when determining the frame rate that should be
used for acquisition.

Give me a call or feel free to email me to discuss your application in
greater detail.

Thanks,
Jim Haley

******************************
Jim Haley
Applications Engineer
MVIA, Inc.
125 Sherwood Drive
Monaca, PA 15061
voice: (724) 728-7493
fax: (412) 291-1709
e-mail: haley-at-mvia.com
webpage: http://www.mvia.com
******************************

Damian Neuberger wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hi Listers,
}
} I have limited experience with video recording so I may not be giving enough
} information; also limited for other reasons. I am in the market for a high
} speed video camera and would like some mfg recomendations. We would like to
} record a process that takes about 1 second and view an area that is about
} 1cm square and have a depth of field about 1cm. In other words, the process
} takes place within a one cubic cm volume. Also want a camera that can
} record without using strobe illumination. Reasonably high resolution (for
} HS Video) 800 by 600 would be nice but probably way too expensive. Maybe 2k
} frames per second? Would also perfer something that is "somewhat" portable.
} Definitely plan to try it out before purchase.
}
} Any recomendations and information in selecting such an instrument would be
} very appreciated.
}
} Damian Neuberger
} Baxter Healthcare Corp.
} damian_neuberger-at-baxter.com


From daemon Wed Oct 9 11:02:36 2002



From: ENGR. JOE UDAH :      joeudah123-at-dpr-nig.com
Date: Wed, 9 Oct 2002 16:49:06 +0100
Subject: DPR

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



DEPARTMENT OF PETROLEUM RESOURCES
PLOT 225 KOFO ABAYOMI STREET VICTORIA ISLAND,LAGOS, NIGERIA.
DIRECT TEL: 234 1 7591519 / FAX: 234 1 7590904
ATTENTION: THE PRESIDENT/C.E.O
RE: URGENT & CONFIDENTIAL BUSINESS PROPOSAL

Dear Sir,

I am ENGR. JOE UDAH ,member committee of the above department
Terms of Reference
My term of reference involves the award of contracts to multinational companies.
My office is saddled with the responsibility of contract award, screening, categorization and prioritization of projects embarked upon by Department of Petroleum Resources (DPR) as well as feasibility studies for selected projects and supervising the project consultants involved. A breakdown of the fiscal expenditure by this office as at the end of last fiscal quarter of 2000 indicates that DPR paid out a whooping sum of US$736M(Seven Hundred And Thirty Six Million, United States Dollars) to successful contract beneficiaries. The DPR is now compiling beneficiaries to be paid for the third Quarter of 2002.
The crux of this letter is that the finance/contract department of the DPR deliberately over –invoiced the contract value of the various contracts awarded. In the course of disbursements, this department has been able to accumulate the sum of US$38.2M(Thirty-eight Million, two hundred Thousand U.S Dollars) as the over-invoiced sum. This money is currently in a suspense account of the DPR account with the Debt Reconciliation Committee (DRC). We now seek to process the transfer of this fund officially as contract payment to you as a foreign contractor, who will be fronting for us as the beneficiary of the fund. In this way we can facilitate these funds into your nominated account for possible investment abroad. We are not allowed as a matter of government policy to operate any foreign account to transfer this fund into.
However, for your involvement in assisting us with this transfer into your nominated account we have evolved a sharing formula as follows:
(1) 20% for you as the foreign partner
(2) 75% for I and my colleagues

(3) 5% will be set aside to defray all incidental expenses both Locally and Internationally during the course of this transaction.

We shall be relying on your advice as regard investment of our share in any business in your country. Be informed that this business is genuine and 100% safe considering the high-power government officials involved. Send your private fax/telephone numbers. Upon your response we shall provide you with further information on the procedures. Feel free to send response by Fax or TEL; expecting your response urgently. All enquiries should be directed to the undersigned by FAX OR PHONE.
Looking forward to a good business relationship with you.

Sincerely,
ENGR. JOE UDAH.




From daemon Wed Oct 9 11:26:26 2002



From: Ann-Fook Yang :      yanga-at-agr.gc.ca
Date: Wed, 09 Oct 2002 11:05:53 -0400
Subject: ESEM for untreated bacteria

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I have been asked to provide a protocol for observing untreated bacteria in ESEM. Can anyone help?


AnnFook Yang
EM Unit,
Eastern Cereal and Oilseed Research Centre,
Room 2091, Bldg. 20,
Central Experimental Farm,
Ottawa, Ontario
Canada K1A 0C6

Tel: 1-613-759-1638
Fax: 1-613-759-1701

e-mail: yanga-at-em.agr.ca

} } } Letizia Anello {anello-at-ibs.pa.cnr.it} 10/03/02 08:50AM } } }
Dear Ann Fook,

I am a molecular biologist and have no notion on
electron microscopy and technical details; so I have collaboration with
a chemical engineer of Palermo University. We have tried to examine
unsuccessfully cultured bacteria directly under low vacuum SEM and ESEM
(can you tell me the enlargement and pressure value to use?). I'd prefer
observe in ESEM, completely untreated bacterial samples. Can you send me
a detail protocol? Can I have any photo of yours of bacterial samples in
SEM and ESEM? Thanks, bye.

Letizia

=====================================







From daemon Wed Oct 9 11:26:26 2002



From: Ted Pella :      ted_pella-at-tedpella.com
Date: Wed, 9 Oct 2002 09:20:16 -0700
Subject: Cressington service in the USA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Listservers:

Following through on a message posted on the Listserver yesterday, we are
clarifying service questions regarding
Cressington equipment as well as OEM Cressington equipment under the PELCO
name.

Please be assured that Cressington vacuum equipment service and parts for
the Americas are available from Ted Pella, Inc. and PELCO International.

This includes versions of the 108 series (plain, /A, /C, /SE), 208 high
resolution series (/HR, /C), 308 series (plain, /R)
and the older versions under the PELCO CC7A, SC-6, SC-7 names.

Our specialist for operation and other questions for the Cressington
equipment can be reached through our office.

You'll find us at 1-800-237-3526, web or e-mail.

Ted Pella




From daemon Wed Oct 9 12:55:03 2002



From: Jay Campbell :      jmcampbe-at-facstaff.wisc.edu
Date: Wed, 09 Oct 2002 13:38:48 -0500
Subject: Re: Fwd: CLAIMS/INVESTMENTS FROM HIS HIGHNESS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



------- Forwarded message follows -------
} From: Marcia A. Miller, Ph.D. ABMM {Marcia-at-dbts.uicomp.uic.edu}
To: Microscopy-at-MSA.Microscopy.Com


Spam? It sounded like a pretty good deal to me. I sent His Highness
my contact info along with bank account and credit card numbers. I
can't wait to start pulling down those 10% commissions!
Jay

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--


From daemon Wed Oct 9 13:50:30 2002



From: Shanling Shi :      Shanling.Shi-at-unilever.com
Date: Wed, 9 Oct 2002 14:42:45 -0400
Subject: Cressington service in the USA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


What about Cressington freeze fracture Unit? I was told that Ted Pella could
not yet do the service for freeze fracture in USA. Is it still true? Thanks.

Shanling

Shanling Shi
Advanced Imaging & Measurement
Unilever Research US
45 River Road
Edgewater, NJ 07020
201-840-2340
Shanling.Shi-at-unilever.com


-----Original Message-----
} From: Ted Pella [SMTP:ted_pella-at-tedpella.com]
Sent: Wednesday, October 09, 2002 12:20 PM
To: microscopy-at-sparc5.microscopy.com



Listservers:

Following through on a message posted on the Listserver yesterday, we are
clarifying service questions regarding
Cressington equipment as well as OEM Cressington equipment under the PELCO
name.

Please be assured that Cressington vacuum equipment service and parts for
the Americas are available from Ted Pella, Inc. and PELCO International.

This includes versions of the 108 series (plain, /A, /C, /SE), 208 high
resolution series (/HR, /C), 308 series (plain, /R)
and the older versions under the PELCO CC7A, SC-6, SC-7 names.

Our specialist for operation and other questions for the Cressington
equipment can be reached through our office.

You'll find us at 1-800-237-3526, web or e-mail.

Ted Pella






From daemon Wed Oct 9 14:14:12 2002



From: Gary Gaugler :      gary-at-gaugler.com
Date: Wed, 09 Oct 2002 12:07:03 -0700
Subject: Re: Fwd: CLAIMS/INVESTMENTS FROM HIS HIGHNESS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Great...here's another one then that you should surely
check out:

MRS.THADDEE KABILA" {bhg777-at-tur.net}

She is:

I AM FORTY-NINE YEARS OLD WIDOW (THE THIRD WIFE) OF THE LATE PRESIDENT
LAURENT KABILA OF THE DEMOCRATIC REPUBLIC OF CONGO

I assume she means that he was late for dinner because
I don't recall the Congo having a president. By the way,
tur.net is in Istanbul, Turkey. The poor widow claims to
be in the U.S. with lots of money.

Same old scams......

gary


At 11:38 AM 10/9/2002, you wrote:
} Spam? It sounded like a pretty good deal to me. I sent His Highness my
} contact info along with bank account and credit card numbers. I can't
} wait to start pulling down those 10% commissions!
} Jay
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Wed Oct 9 14:14:13 2002



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Thu, 10 Oct 2002 08:01:26 +1200
Subject: JEOL 840A airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Ken


}
} The second was not so bad, but also having to do with the airlock. If
} the airlock is left open (not pumped) most of the time then the
} airlock vent valve (a small solenoid valve) gets quite hot after a
} while and remains hot (sometimes for years on end). This hardens the
} rubber insert in the plunger so that it will no longer seal properly.
} JEOL operates their airlock on a timer rather than from Pirrani
} readings, so when it times out it says you can transfer the specimen.
} This can lead to a considerable gas pulse and some minor
} backstreaming from the DP before the high vacuum valve closes.
}

Yeah, my 840A has that condition. I put three manual ball-type valves
between the airlock and a spare rotary pump, with a T/C vacuum
gauge in there too, as it seemed like I was going to have to do a lot of
dismantling to fix that valve. The manual system works so well, I find
that 10 minutes sucking on samples before insertion really reduces the
degradation of chamber vacuum on opening the gate valve, that I've
left it like that.

But maybe I should fix the vent valve anyway. Anybody got any tips on
how that get at the vent valve with the minimum of collateral
disassembly?

cheers

rtch

Ritchie Sims Phone : 64 9 3737599 ext 7713
Department of Geology Fax : 64 9 3737435
The University of Auckland email : r.sims-at-auckland.ac.nz
Private Bag 92019
Auckland
New Zealand



From daemon Wed Oct 9 16:13:48 2002



From: Beavers, Roy :      rbeavers-at-post.cis.smu.edu
Date: Wed, 9 Oct 2002 15:57:22 -0500
Subject: View Petrographic Slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Group,


Looking for a method of projecting a petrographic thin section (27X46mm) onto a large screen to be viewed by students. Have tried to sandwich the slide between two pieces of polarized film and display on a microfiche reader and that works pretty well but would like a larger brighter image.

Have tried a digital camera on a petro microscope and again works well but area of view is limited.

Any ideas?

Roy Beavers
Southern Methodist University
Department of Geological Sciences
P.O. Box 750395
Dallas, Tx. 75275
Voice: 214-768-2756
Fax: 214-768-2701
Email: rbeavers-at-mail.smu.edu



From daemon Wed Oct 9 17:43:48 2002



From: qualityimages :      qualityimages-at-netrax.net
Date: Wed, 09 Oct 2002 18:30:03 -0400
Subject: Re: JEOL 840A airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Ritch,
The valve isn't that hard to get to. JEOL uses all those nice
easy-to-assemble connectors (can't say quick connect 'cause they're not
QF or KF) so after cooling the system down (although if you're really
careful, I think I did it with a hot system - look at it very carefully
before proceding hot), you can detatch a pair of solenoid valves from
the roughing manifold. DO NOT separate the valves from the pipe
fittings. JEOL epoxies them together and you may destroy something
trying to separate them. Just move the assembly out to where it is a
little easier to work on and disassemble the valve and remove the
plunger. I bought a new valve ($100.00 US)for one because I was only 20
miles from their US headquarters and just replaced the plunger, but for
the second one (which hasn't been installed yet) I brought the old
plunger home and dug out all the old, hard rubber. Then I took a cork
borer (I forget what size, but a snug fit) and cut a plug out of a
rubber stopper, cut it to the appropriate length, stuck some "Goop[" (an
everything, thick adhesive) into the hole in the plunger and rammed the
plug in so that is was level with the end of the plunger and let it set.
I'm expecting it to give me a good seal and a long life.


Ken Converse
owner
Quality Images
Delta, PA


Ritchie Sims wrote:

} Hi Ken
}
}
}
}
} } The second was not so bad, but also having to do with the airlock. If
} } the airlock is left open (not pumped) most of the time then the
} } airlock vent valve (a small solenoid valve) gets quite hot after a
} } while and remains hot (sometimes for years on end). This hardens the
} } rubber insert in the plunger so that it will no longer seal properly.
} } JEOL operates their airlock on a timer rather than from Pirrani
} } readings, so when it times out it says you can transfer the specimen.
} } This can lead to a considerable gas pulse and some minor
} } backstreaming from the DP before the high vacuum valve closes.
} }
} }
} }
}
} Yeah, my 840A has that condition. I put three manual ball-type valves
} between the airlock and a spare rotary pump, with a T/C vacuum
} gauge in there too, as it seemed like I was going to have to do a lot of
} dismantling to fix that valve. The manual system works so well, I find
} that 10 minutes sucking on samples before insertion really reduces the
} degradation of chamber vacuum on opening the gate valve, that I've
} left it like that.
}
} But maybe I should fix the vent valve anyway. Anybody got any tips on
} how that get at the vent valve with the minimum of collateral
} disassembly?
}
} cheers
}
} rtch
}
} Ritchie Sims Phone : 64 9 3737599 ext 7713
} Department of Geology Fax : 64 9 3737435
} The University of Auckland email : r.sims-at-auckland.ac.nz
} Private Bag 92019
} Auckland
} New Zealand
}
}
}
}





From daemon Wed Oct 9 20:48:41 2002



From: Allen Sampson :      ars-at-sem.com
Date: Wed, 9 Oct 2002 20:42:48 -0700
Subject: RE: JEOL 840A airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I haven't run across that problem myself, but sounds like a situation where
a simple circuit modification would be the best route.

In order to activate the valve, a small power transistor would be used to
ground one end of the coil while the other end is always attached to the
power supply. A small voltage (5V) would activate the transistor (ground
the coil and open the valve), 0V would deactivate it. Send that signal
instead to a simple multivibrator one-shot circuit set up to produce a
pulse with a duration of perhaps a minute, and send that to the transistor.
The airlock vents in just a few seconds, so it will always vent OK, but
would shut a minute later, thus limiting the time it is left activated.
Properly designed, if the excitation circuit goes low, the circuit would
automatically reset, in case the operator hits the button again and starts
the airlock roughing before the timer times out.

I don't have those schematics in front of me, but it's a pretty standard
way of running a solenoid valve. It should be easy to find someone to make
such a simple modification.


Allen R. Sampson
Advanced Research Systems
317 North 4th. Street
St. Charles, Illinois 60174

phone (630) 513-7093 fax (630) 513-7092 http://www.sem.com


On Wednesday, October 09, 2002 1:01 PM, Ritchie Sims
[SMTP:r.sims-at-auckland.ac.nz] wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi Ken
}
}
} }
} } The second was not so bad, but also having to do with the airlock. If
} } the airlock is left open (not pumped) most of the time then the
} } airlock vent valve (a small solenoid valve) gets quite hot after a
} } while and remains hot (sometimes for years on end). This hardens the
} } rubber insert in the plunger so that it will no longer seal properly.
} } JEOL operates their airlock on a timer rather than from Pirrani
} } readings, so when it times out it says you can transfer the specimen.
} } This can lead to a considerable gas pulse and some minor
} } backstreaming from the DP before the high vacuum valve closes.
} }
}
} Yeah, my 840A has that condition. I put three manual ball-type valves
} between the airlock and a spare rotary pump, with a T/C vacuum
} gauge in there too, as it seemed like I was going to have to do a lot of
} dismantling to fix that valve. The manual system works so well, I find
} that 10 minutes sucking on samples before insertion really reduces the
} degradation of chamber vacuum on opening the gate valve, that I've
} left it like that.
}
} But maybe I should fix the vent valve anyway. Anybody got any tips on
} how that get at the vent valve with the minimum of collateral
} disassembly?
}
} cheers
}
} rtch
}
} Ritchie Sims Phone : 64 9 3737599 ext 7713
} Department of Geology Fax : 64 9 3737435
} The University of Auckland email : r.sims-at-auckland.ac.nz
} Private Bag 92019
} Auckland
} New Zealand
}
}



From daemon Wed Oct 9 21:55:15 2002



From: Philip Oshel :      peoshel-at-facstaff.wisc.edu
Date: Wed, 09 Oct 2002 21:43:39 -0500
Subject: Re: View Petrographic Slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


How about a projecting microscope? Built somewhat like an inverted on
a really tall stand, and no tube. The objective projects the image
onto a surface (like a sheet of paper) 30 - 50 cm or more below the
stage. Great for drawing specimens, and should work for this purpose,
although you'll need to darken the room.
Also cheap, and old tech, so there should be plenty of used units around.
Phil

} Group,
}
}
} Looking for a method of projecting a petrographic thin section
} (27X46mm) onto a large screen to be viewed by students. Have tried
} to sandwich the slide between two pieces of polarized film and
} display on a microfiche reader and that works pretty well but would
} like a larger brighter image.
}
} Have tried a digital camera on a petro microscope and again works
} well but area of view is limited.
}
} Any ideas?
}
} Roy Beavers
} Southern Methodist University
} Department of Geological Sciences
} P.O. Box 750395
} Dallas, Tx. 75275
} Voice: 214-768-2756
} Fax: 214-768-2701
} Email: rbeavers-at-mail.smu.edu
}

--
}}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
Philip Oshel
Supervisor, AMFSC and BBPIC microscopy facilities
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706 - 1284
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)


From daemon Wed Oct 9 22:37:38 2002



From: John Twilley :      jtwilley-at-sprynet.com
Date: Wed, 09 Oct 2002 23:36:53 -0400
Subject: Re: View Petrographic Slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Roy,

Get a plastic photo slide frame; trim away the ridges on one side that are intended to hold the 35mm film in place so that the petro slide can stick out without being shortened; cut a piece of polarizing film to fit inside the frame behind the petro slide and close it with a piece of tape.

Next put a piece of polarizing film at right angles to the first over the front of a projector lens; drop the slide in its mount into the projector with the polarizer behind, and away you go. If it is an autofocus projector you may need to disable the autofocus drive so as to prevent it from
focusing on the rear surface of the polarizer instead of the thin section.

This is also a means of photographing a very coarse grained rock between crossed polars when the field of view would exceed that of your microscope. Simply set up the camera over the projector and photograph the image on the screen. This works better than using a slide scanner and two pieces
of polarizing film, in many cases, because it keeps the polarizers out of the focal plane so that their imperfections are not superimposed on the thin section.

John Twilley
Conservation Scientist

Beavers, Roy wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Group,
}
} Looking for a method of projecting a petrographic thin section (27X46mm) onto a large screen to be viewed by students. Have tried to sandwich the slide between two pieces of polarized film and display on a microfiche reader and that works pretty well but would like a larger brighter image.
}
} Have tried a digital camera on a petro microscope and again works well but area of view is limited.
}
} Any ideas?
}
} Roy Beavers
} Southern Methodist University
} Department of Geological Sciences
} P.O. Box 750395
} Dallas, Tx. 75275
} Voice: 214-768-2756
} Fax: 214-768-2701
} Email: rbeavers-at-mail.smu.edu





From daemon Thu Oct 10 00:48:02 2002



From: Arthur Day :      ard-at-ansto.gov.au
Date: Thu, 10 Oct 2002 15:37:51 +1000
Subject: Re: JEOL 840A airlock

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Rtch,

When I did it on our 6400 (which in our case had exactly the same
valve as on the 840 we used to have), I actually bought a whole
replacement solenoid valve in preparation. However, as you've
implied, it is a bit of a mongrel of a task to chop the whole thing
out so in the end all I did was pull the solenoid plunger out of the
existing valve (with the little bit of heat-damaged rubber on its
end) and replaced it with just the new plunger and rubber bit from
the new valve. That way the whole task was a lot less invasive and
offensive to the instrument than it otherwise could have been....

So, see if you can get to just the plunger and then maybe you could
go one better and figure out a way to restore/replace the tiny little
bit of sealing rubber on its end.....

Cheers

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





Arthur Day, Electron Microscope Unit Phone: 61-2-9717-3457
Ansto Materials Division Fax: 61-2-9543-7179
PMB 1, Menai (Sydney), NSW, 2234 Email: ard-at-ansto.gov.au
Australia www: http://www.ansto.gov.au/


From daemon Thu Oct 10 01:54:11 2002



From: Malcolm Roberts :      m.roberts-at-ru.ac.za
Date: Thu, 10 Oct 2002 08:45:49 +0200
Subject: Re: View Petrographic Slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Roy,
Have you thought about acquiring one of these things called a "petroscope"? It's a bit like a glorified microfiche reader but designed for looking at thin-sections with lots of people about. It has a screen about the same size as a moderate TV, different magnifications and PPL and XPL
options. The company that produces these now has a digital camera attachment option available.
Just like to add that I have no interest in this company or whoever they are. And even though we have a petroscope in the department, it is rarely used for reasons that are uncertain.
Cheers,
Malc.

"Beavers, Roy" wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Group,
}
} Looking for a method of projecting a petrographic thin section (27X46mm) onto a large screen to be viewed by students. Have tried to sandwich the slide between two pieces of polarized film and display on a microfiche reader and that works pretty well but would like a larger brighter image.
}
} Have tried a digital camera on a petro microscope and again works well but area of view is limited.
}
} Any ideas?
}
} Roy Beavers
} Southern Methodist University
} Department of Geological Sciences
} P.O. Box 750395
} Dallas, Tx. 75275
} Voice: 214-768-2756
} Fax: 214-768-2701
} Email: rbeavers-at-mail.smu.edu
}

--
Dr MP Roberts Phone: [+27](0)46 603 8313 (work)
Dept of Geology [+27] (0)46 6361197 (home)
Rhodes University Fax: [+27](0)46 622 9715
6140 Grahamstown Cell: 083 4060 262
SOUTH AFRICA e-mail: m.roberts-at-ru.ac.za




From daemon Thu Oct 10 02:22:27 2002



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Thu, 10 Oct 2002 09:14:17 +0200
Subject: Re: View Petrographic Slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I remember Leitz had such a device, a slide projector modified to put
polariser, rotating analyser, and rotating mount (microscope like) for the
slide. We made a copy, modifying a old 2.1/4x2.1/4 SFOM slide projector.
We put the analyser on the objectif, to be able to rotate it. It's a bit
more difficult to make a rotary device fr the slide, but on such a
projector you have place enough to put a rotating table from an old
microscope. Our worked well. The 2.1/4x2.1/4 slide give you more space to
put the petrographic slide than the 24x36, and the objectives will give a
very good large image on the screen. You can than project it on a frosted
glass or a sheet of tracing paper to take a picture with a digital camera
and to save it as digital document.

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess
67037 Strasbourg CEDEX
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr

On Wed, 9 Oct 2002, Beavers, Roy wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Group,
}
}
} Looking for a method of projecting a petrographic thin section (27X46mm) onto a large screen to be viewed by students. Have tried to sandwich the slide between two pieces of polarized film and display on a microfiche reader and that works pretty well but would like a larger brighter image.
}
} Have tried a digital camera on a petro microscope and again works well but area of view is limited.
}
} Any ideas?
}
} Roy Beavers
} Southern Methodist University
} Department of Geological Sciences
} P.O. Box 750395
} Dallas, Tx. 75275
} Voice: 214-768-2756
} Fax: 214-768-2701
} Email: rbeavers-at-mail.smu.edu
}
}



From daemon Thu Oct 10 05:28:35 2002



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Thu, 10 Oct 2002 11:19:37 +0100 (GMT Daylight Time)
Subject: Re: ESEM for untreated bacteria

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


In the UK we have to make a safety assessment re using
unfixed bacteria in the ESEM. For instance I insist that
clinical urinary catheters are fixed.

Unfixed bacteria.
The magic of ESEM is that you can just have a go and then
look round for improvements! I always start with the
easiest method.

If they are in suspension I would pellet them gently and
rinse in distilled water and spin gently. Put the pellet
on the Peltier stage at 5 degree C with a little water to
cover it. Pump down at 7.5 Torr to avoid dehydration then
slowly reduce pressure to remove water and expose bacteria.

Disclaimer: ESEM images of bacteria have a different
quality to those that have been dried and gold coated; so
warn your user to avoid disappointment.

Let us know if this works for you!

Dave


On Wed, 09 Oct 2002 11:05:53 -0400 Ann-Fook Yang
{yanga-at-agr.gc.ca} wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I have been asked to provide a protocol for observing untreated bacteria in ESEM. Can anyone help?
}
}
} AnnFook Yang
} EM Unit,
} Eastern Cereal and Oilseed Research Centre,
} Room 2091, Bldg. 20,
} Central Experimental Farm,
} Ottawa, Ontario
} Canada K1A 0C6
}
} Tel: 1-613-759-1638
} Fax: 1-613-759-1701
}
} e-mail: yanga-at-em.agr.ca
}
} } } } Letizia Anello {anello-at-ibs.pa.cnr.it} 10/03/02 08:50AM } } }
} Dear Ann Fook,
}
} I am a molecular biologist and have no notion on
} electron microscopy and technical details; so I have collaboration with
} a chemical engineer of Palermo University. We have tried to examine
} unsuccessfully cultured bacteria directly under low vacuum SEM and ESEM
} (can you tell me the enlargement and pressure value to use?). I'd prefer
} observe in ESEM, completely untreated bacterial samples. Can you send me
} a detail protocol? Can I have any photo of yours of bacterial samples in
} SEM and ESEM? Thanks, bye.
}
} Letizia
}
} =====================================
}
}
}
}
}
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



From daemon Thu Oct 10 05:49:18 2002



From: mohammed y abdulrawoof / f40z006 75760 :      mdyousuf-at-KSU.EDU.SA
Date: Thu, 10 Oct 2002 13:35:19 -0300 (GMT)
Subject: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear listers,
Please advise as to the usefulness/advantage of doing enblock staining of
biological tissues as against doing heavy-metal staining of grids with
thin sections. The tissues that I am about to process are for diagnosis
(pathology).

Thanks
M. Yousuf Abdul-Rawoof



From daemon Thu Oct 10 06:07:33 2002



From: Tim E. Harper :      tim-at-cmp-cientifica.com
Date: Thu, 10 Oct 2002 13:05:28 +0200
Subject: EU Spending on Nanotechnology double that of US?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

The European NanoBusiness Association recently produced a report on EU spending on nanotechnology - with the conclusions being that it is currently higher than the US, and very focused on applications and solutions. For anyone interested, it is a free download from www.nanoeurope.org

Best

Tim

*****************************************************************
Empowering business to make rational decisions about nanotechnology

CMP Cientifica
http://www.cmp-cientifica.com

Tim E. Harper - CEO
+34 91 640 71 85 (direct)
+ 34 91 640 71 86 (fax)
} From USA (877) 295-4480 (phone/fax)

CV: http://www.google.com/search?q=tim+%2B+harper+%2B+nanotechnology

The X Report - The crossroads of info- bio- & nano- tech
www.cmp-cientifica.com/x

The European Nanobusiness Association
www.nanoeurope.org

NanotechWeb - News and Information
www.nanotechweb.org

Nanoelectronics Reseach - The Phantoms Network
www.phantomsnet.com




From daemon Thu Oct 10 07:30:22 2002



From: John.Catino-at-mineralstech.com
Date: Thu, 10 Oct 2002 08:21:57 -0400
Subject: Job Opening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


ANALYTICAL INVESTIGATOR


Minerals Technologies Inc., an international resource and technology based
organization, currently has an opening for a Analytical Investigator
Easton, PA facility

The Analytical Investigator is responsible for developing and completing
analytical investigations of samples, maintaining technical mastery and
awareness for microscopy techniques, assisting in the accreditation to ISO
17025, and completing administrative duties. The primary duty is
customer-oriented problem solving using optical and electron microscopy and
microchemical analysis in a team environment. FT-IR and GC-MS techniques
may also be utilized. The Analytical Investigator completes work of a
varied nature and is responsible for making and implementing most of the
decisions relevant to their area of responsibility.

Requirements include a Bachelor's degree in the sciences preferably
chemistry or mineralogy and a minimum of 5 years microscopy experience in a
service environment. Experience in computer assisted microscopy is highly
desirable. All candidates must possess good oral and written communication
skills, computer skills, ability to interact with a variety of people,
willingness to work in a team environment, to work extended hours when
necessary and ability to operate instrumentation and to manipulate
equipment and materials weighing 30-50 pounds. Previous experience with
LIMS, XRF or ICP spectroscopy is desirable.

MTI offers a competitive compensation and benefits package. Please submit
resume and cover letter with salary history to:

Bernadette Palumbo
Human Resources Development Manager
Minerals Technologies Inc.
640 N. 13th Street
Easton, PA 18042
Fax: 610-250-3210
Email: bpalumbo-at-mineralstech.com
No phone calls, please. Only those candidates meeting our requirements will
be contacted. Equal Opportunity Employer M/F/D/V.



John Catino
Minerals Technologies, Inc.
640 N. 13th Street
Easton, PA 18042
Phone: 610-250-3363
Fax: 610-250-3206
e-mail: john.catino-at-mineralstech.com



**********************************************************************
This email and any files transmitted with it are confidential and
intended solely for the use of the individual or entity to whom they
are addressed. If you have received this email in error please notify
the system manager.

This footnote also confirms that this email message has been swept by
MIMEsweeper for the presence of computer viruses.
www.mimesweeper.com
**********************************************************************



From daemon Thu Oct 10 07:32:40 2002



From: Scott Whittaker :      Whittaker.scott-at-nmnh.si.edu
Date: Thu, 10 Oct 2002 08:26:18 -0400
Subject: Re: ESEM for untreated bacteria

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


First of all you MUST consider the health ramifications of untreated
pathogens in the chamber. They can contaminate the whole instrument and be
aerosolized into the room. I assume this has been considered and only
provide for the sake of completeness since this is going out to the whole
list.

Instrumentation wise as far as I know this can only be done with the
FEI/Philips/Electroscan instruments capable of condensing water in the
chamber. Not sure of the pressure ranges for the other manufacturers, but at
2 Torr chamber pressure you would freeze the sample before being able to
condense water or maintain the hydrated state. A field emission instrument
would give best results, but I have a LaB6 instrument and imaged similar
quite nicely on the Peltier stage cooled to 1C and 5 Torr vapor pressure.

Procedure wise there are several issues to overcome. First is the culture
itself. Is it a broth or plated. Both can be imaged each presenting
different issues. Precipitation of salts is an issue for both. For a plated
sample simply punch a thin core out and place flat on a petri. Trim excess
agar away. Surround with a drop of distilled water without letting any get
on top. This will remove some of the salts that could punch through the
colony. This step can be omitted if you are real careful with your vacuum
conditions. Plated samples may also have a polysaccharide film covering them
preventing imaging of the organism.

Broth cultures again have all kinds of problems with salts and gunk coating
the organism. They should be centrifuged loosely down, and washed with
distilled water. Place a drop of resuspended organism directly on the
Peltier stage and carefully evaporate the water in the chamber.

Both are still highly susceptible to the vacuum, reduced though it is, and
beam damage so image quickly and with lower voltages. This is really why a
FE-ESEM would be recommended.

Good luck

Scott Whittaker
Laboratories of Analytical Biology
Smithsonian Institution
National Museum of Natural History
PO Box 37012 MRC104
Washington DC 20013-7012
202-357-1651


} } } "Ann-Fook Yang" {yanga-at-agr.gc.ca} 10/09/02 11:05AM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


I have been asked to provide a protocol for observing untreated bacteria in
ESEM. Can anyone help?


AnnFook Yang
EM Unit,
Eastern Cereal and Oilseed Research Centre,
Room 2091, Bldg. 20,
Central Experimental Farm,
Ottawa, Ontario
Canada K1A 0C6

Tel: 1-613-759-1638
Fax: 1-613-759-1701

e-mail: yanga-at-em.agr.ca

} } } Letizia Anello {anello-at-ibs.pa.cnr.it} 10/03/02 08:50AM } } }
Dear Ann Fook,

I am a molecular biologist and have no notion on
electron microscopy and technical details; so I have collaboration with
a chemical engineer of Palermo University. We have tried to examine
unsuccessfully cultured bacteria directly under low vacuum SEM and ESEM
(can you tell me the enlargement and pressure value to use?). I'd prefer
observe in ESEM, completely untreated bacterial samples. Can you send me
a detail protocol? Can I have any photo of yours of bacterial samples in
SEM and ESEM? Thanks, bye.

Letizia

=====================================








From daemon Thu Oct 10 07:36:47 2002



From: Kestutis Smalinskas :      smalinskas-at-yahoo.com
Date: Thu, 10 Oct 2002 05:30:24 -0700 (PDT)
Subject: Re: Fwd: CLAIMS/INVESTMENTS FROM HIS HIGHNESS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Gary, I simply forwarded the e-mail from HIS HIGHNESS
the Emir of Bahrain to Joe Udah from the Department of
Petroleum Resources in Nigeria. Let them work it out
amongst themselves!

Stu Smalinskas

__________________________________________________
Do you Yahoo!?
Faith Hill - Exclusive Performances, Videos & More
http://faith.yahoo.com


From daemon Thu Oct 10 10:18:20 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Thu, 10 Oct 2002 11:07:40 -0400
Subject: DPR

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Mr. Udah;

I will refer your request to a high powered government official in the
Federal Prosecutor's Office in Newark, New Jersey.

Sincerely,

P. Tomic

-----Original Message-----
} From: ENGR. JOE UDAH [mailto:joeudah123-at-dpr-nig.com]
Sent: Wednesday, October 09, 2002 11:49 AM
To: microscopy-at-sparc5.microscopy.com



DEPARTMENT OF PETROLEUM RESOURCES
PLOT 225 KOFO ABAYOMI STREET VICTORIA ISLAND,LAGOS, NIGERIA.
DIRECT TEL: 234 1 7591519 / FAX: 234 1 7590904
ATTENTION: THE PRESIDENT/C.E.O
RE: URGENT & CONFIDENTIAL BUSINESS PROPOSAL

Dear Sir,

I am ENGR. JOE UDAH ,member committee of the above department
Terms of Reference
My term of reference involves the award of contracts to multinational
companies.
My office is saddled with the responsibility of contract award, screening,
categorization and prioritization of projects embarked upon by Department of
Petroleum Resources (DPR) as well as feasibility studies for selected
projects and supervising the project consultants involved. A breakdown of
the fiscal expenditure by this office as at the end of last fiscal quarter
of 2000 indicates that DPR paid out a whooping sum of US$736M(Seven Hundred
And Thirty Six Million, United States Dollars) to successful contract
beneficiaries. The DPR is now compiling beneficiaries to be paid for the
third Quarter of 2002.
The crux of this letter is that the finance/contract department of the DPR
deliberately over -invoiced the contract value of the various contracts
awarded. In the course of disbursements, this department has been able to
accumulate the sum of US$38.2M(Thirty-eight Million, two hundred Thousand
U.S Dollars) as the over-invoiced sum. This money is currently in a suspense
account of the DPR account with the Debt Reconciliation Committee (DRC). We
now seek to process the transfer of this fund officially as contract payment
to you as a foreign contractor, who will be fronting for us as the
beneficiary of the fund. In this way we can facilitate these funds into your
nominated account for possible investment abroad. We are not allowed as a
matter of government policy to operate any foreign account to transfer this
fund into.
However, for your involvement in assisting us with this transfer into your
nominated account we have evolved a sharing formula as follows:
(1) 20% for you as the foreign partner
(2) 75% for I and my colleagues

(3) 5% will be set aside to defray all incidental expenses both Locally and
Internationally during the course of this transaction.

We shall be relying on your advice as regard investment of our share in any
business in your country. Be informed that this business is genuine and 100%
safe considering the high-power government officials involved. Send your
private fax/telephone numbers. Upon your response we shall provide you with
further information on the procedures. Feel free to send response by Fax or
TEL; expecting your response urgently. All enquiries should be directed to
the undersigned by FAX OR PHONE.
Looking forward to a good business relationship with you.

Sincerely,
ENGR. JOE UDAH.




From daemon Thu Oct 10 11:01:04 2002



From: saram-at-duke.edu
Date: Thu, 10 Oct 2002 11:47:20 -0400 (EDT)
Subject: Re: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Uranyl acetate stabilizes lipids, membranes, and nucleic acids when used
as a fixative. Studies have shown less lipid extraction by the organic
solvents after UA fixation. It also contributes somewhat to contrast,
though not quite as much as osmium (i.e., if you leave out Os, and use UA
in fixation, there will be less contrast than if you use Os and not UA).
UA is generally used after osmication for additional stabilization and
contrast. It can be used in veranal acetate buffer, aqueous solutions,
or in alcoholic solutions.

We routinely use UV en bloc in pathology and virology samples EXCEPT in
cases where one needs to see glycogen deposits (e.g., glycogen storage
diseases). Thus, we never use UA in cases of heart, skeletal muscle, and
Ewing's sarcoma (if Ewing's is suspected).

UA and lead, used as post stains, on sections add a general contrast to
structures present (versus cytosol or plastic). In this case, UA does
not contribute to fixation or stabilization because the structures are
already embedded in resin.

Hope this helps.

Sara Miller


On Thu, 10 Oct 2002, mohammed y abdulrawoof / f40z006 75760 wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear listers,
} Please advise as to the usefulness/advantage of doing enblock staining of
} biological tissues as against doing heavy-metal staining of grids with
} thin sections. The tissues that I am about to process are for diagnosis
} (pathology).
}
} Thanks
} M. Yousuf Abdul-Rawoof
}
}
}
}

Sara E. Miller, Ph. D.
P. O. Box 3712
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-3265



From daemon Thu Oct 10 11:53:04 2002



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Thu, 10 Oct 2002 12:40:14 -0500
Subject: ESEM vs. gold coated SEM images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

David Patton wrote:
==========================================================
Disclaimer: ESEM images of bacteria have a different quality to those that
} have been dried and gold coated; so warn your user to avoid
disappointment.
===========================================================

Are you comparing ESEM images with a) air dried and gold coated or b)
critical point dried and gold coated?

It would be obvious that comparing air dried would indeed show a difference
but if you really were comparing it to critical point dried samples, could
you elaborate a bit on the nature of those differences? And would you
have any comment about differences in contrast comparing ESEM vs. gold
coated images.


Disclaimer: SPI Supplies offers critical point dryers and sputter coaters so
I guess we would be a bit nervous of any technology that should evolve that
would render obsolete at the same time both CPD units and sputter coaters!

Chuck
============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================


From daemon Thu Oct 10 14:02:07 2002



From: Beavers, Roy :      rbeavers-at-post.cis.smu.edu
Date: Thu, 10 Oct 2002 13:50:15 -0500
Subject: Thanks for Viewer ideas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Group,

Thanks for all of you who offered suggestions on how to project a large view of a petrographic slide.

With the help of my machine shop, I was able to design a petro slide holder that would fit into a standard 35mm slide projector. Once polarized film was taped to the outsides, we were able to project a large room size cross polar image of what students see in their microscopes at low mag.

Again thanks for the great ideas.

Roy Beavers
Southern Methodist University
Department of Geological Sciences
P.O. Box 750395
Dallas, Tx. 75275
Voice: 214-768-2756
Fax: 214-768-2701
Email: rbeavers-at-mail.smu.edu



From daemon Thu Oct 10 14:55:32 2002



From: Tim E. Harper :      tim-at-cmp-cientifica.com
Date: Thu, 10 Oct 2002 21:46:06 +0200
Subject: EU Spending on Nanotechnology double that of US?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

The European NanoBusiness Association recently produced a report on EU
spending on nanotechnology - with the conclusions being that it is
currently higher than the US, and very focused on applications and
solutions. For anyone interested, it is a free download from
www.nanoeurope.org

Best

Tim

*****************************************************************
Empowering business to make rational decisions about nanotechnology

CMP Cientifica
http://www.cmp-cientifica.com

Tim E. Harper - CEO
+34 91 640 71 85 (direct)
+ 34 91 640 71 86 (fax)
} From USA (877) 295-4480 (phone/fax)


The X Report - The crossroads of info- bio- & nano- tech

www.cmp-cientifica.com/x

The European Nanobusiness Association
www.nanoeurope.org

NanotechWeb - News and Information
www.nanotechweb.org

Nanoelectronics Reseach - The Phantoms Network
www.phantomsnet.com






From daemon Thu Oct 10 17:17:32 2002



From: John Shields :      jshields-at-cb.uga.edu
Date: Thu, 10 Oct 2002 18:05:39 -0400
Subject: Job opening - Avigenics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Post-doc/Microscopist

AviGenics, inc. (www.avigenics.com), the world leader in avian transgenesis
has an immediate post-doc or Research Associate position in our Technology
Development group. The appointment will be for one year, with possibility of
renewal.

The successful applicant will join a well funded, dynamic and
multidisciplinary team developing new transgenic technology combining
advanced microscopy methods, such as multiphoton, and cell and embryo
mechanical as well as laser-mediated micromanipulation techniques.

Hands-on experience with confocal and/or multiphoton microscopy is highly
desirable. We are particularly interested in candidates with demonstrated
experience on imaging optically opaque and live samples.

Requirements: preference will be given to candidates with a PhD degree, but
other suitable applicants with proven experience in the areas described
above will also be considered. Must be able to handle multiple projects
simultaneously, keeping accurate, detailed and well organized records, able
to work in a team environment and a have a strong interest in applied
research and technology development.

Salary and benefits including stock options will be commensurate with
experience

For information and submitting applications please contact:

Leandro Christmann DVM, MSc, PhD

Director of Technology Development

AviGenics, inc.

Georgia BioBusiness Center

111 Riverbend Rd.

Athens, GA 30605

Fax: (706) 227-2180

e-mail: christmann-at-avigenics.com




From daemon Thu Oct 10 17:52:53 2002



From: Libby Shaw :      elshaw-at-mit.edu
Date: Thu, 10 Oct 2002 19:10:20 -0400
Subject: Re: those REQUESTS FOR YOUR HELP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Dorota,

Fight this with all you have.

The institution I just quit downgraded all of the lab staff positions. A
microbiologist with no em experience replaced me with 20 years of em
experience only because she lost her job due to downsizing and bumped me out
of my position. The management had recently changed (lowered) all of the
standards/qualifications so this event could take place. The union was
never informed of this move, but that is another story.

A supervisors position should not be part of the union. A smart manager
should realize that you cannot hire good help in this field into a
supervisor position without at least 7-10 years of EM experience.

The minimum education should be at least a BS in a biological field with at
least 5 years experience.

Cheers,

Ed Calomeni



----- Original Message -----
} From: "Dorota Wadowska" {wadowska-at-upei.ca}
To: {microscopy-at-sparc5.microscopy.com}
Sent: Tuesday, October 08, 2002 10:09 AM


You'll find an entertainingly relevant article at
http://www.theatlantic.com/issues/2002/10/working.htm

Libby Shaw




****************************************************************
Elisabeth L. Shaw, Facility Coordinator
Surface and Spectroscopy Labs
Analytical Shared Experimental Facilities
MIT Center for Materials Science and Engineering

Address: MIT Room 13-4149 Tel: 617-253-5045
77 Massachusetts Avenue Email: elshaw-at-mit.edu
Cambridge, MA 02139 Fax: 617-258-6478
http://web.mit.edu/cmse/www/
****************************************************************


From daemon Thu Oct 10 18:17:49 2002



From: msandova-at-nd.edu (by way of Ask-A-Microscopist)
Date: Thu, 10 Oct 2002 18:11:59 -0500
Subject: Ask-A-Microscopist: Whipple Hauser optical piece

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (msandova-at-nd.edu) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Thursday,
October 10, 2002 at 17:40:13
---------------------------------------------------------------------------

Email: msandova-at-nd.edu
Name: Mayra

Organization: U of Notre Dame

Education: Graduate College

Location: Notre Dame, IN

Question: What is and how does one use a Whipple Hauser optical piece?
Thnx!

---------------------------------------------------------------------------


From daemon Thu Oct 10 18:17:48 2002



From: sundeep.bhandari-at-nikon.co.uk (by way of Ask-A-Microscopist)
Date: Thu, 10 Oct 2002 18:11:05 -0500
Subject: Ask-A-Microscopist:epi-fluorescence

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (sundeep.bhandari-at-nikon.co.uk) from on Thursday,
October 10, 2002 at 12:16:30
---------------------------------------------------------------------------

Email: sundeep.bhandari-at-nikon.co.uk
Name: Sundeep Bhandari

Organization: Nikon UK

Education: Graduate College

Location: Kingston, Surrey, England

Question:
I am trying to find information about a technique i heard about a few
days ago, but can't find any information anywhere on the web. The
technique i am researching is 'polarised epi-fluorescence'. I
believe you can use it to determine the surface bound orientation of
fluorophores; i think the
excitation source is linearly polarised and one monitors the
fluorescence intensity as a function of polarisation angle.

Any information or pointers in the right direction would be greatly
appreciated.

kindest rgds
sundeep.

---------------------------------------------------------------------------


From daemon Thu Oct 10 18:21:30 2002



From: Marek Malecki, M.D., Ph.D., Professor :      mmalecki-at-facstaff.wisc.edu
Date: Thu, 10 Oct 2002 18:20:58 -0700
Subject: polishing 1/2 " Cambridge carbon SEM mounts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have attempted to polish /seal / ? carbon 1/2 " carbon SEM
mounts to obtain very smooth surface for HR SEM and be willing to share the
procedure? Ultrafine-grain sand paper , carbon colloid, and card-board did
not generate a surface which is smooth enough.


From daemon Thu Oct 10 20:47:32 2002



From: Lucille A. Giannuzzi :      lag-at-mail.ucf.edu
Date: Thu, 10 Oct 2002 21:38:50 -0400
Subject: FIB Focused Interest Group within MSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


The Microscopy Society of America has recently approved the formation of a
focused ion beam (FIB) focus interest group (FIG). The function of the FIB
FIG will be to organize a symposium for the Microscopy and Microanalysis
annual meeting, to hold an annual business meeting of the FIG, and at the
discretion of the members, hold a workshop or other FIB sponsored event
throughout the year. Individual membership is open to all MSA members at
an additional nominal cost of $10 per year above the annual MSA dues
commencing with the 2003 yearly MSA dues. A healthy membership within the
FIB FIG will ensure that FIB and related topics are an integral part of the
M&M meeting. In addition, members of the FIB FIG will have access to
workshops and programs designed to further the development and applications
of FIB. Vendors are welcome and encouraged to participate in the FIB FIG.
Vendors may join the FIB FIG at a nominal cost of $50 per year.

The first FIB FIG workshop is tentatively set for Tuesday March 18, 2003,
at the University of Central Florida in Orlando during the joint annual
meeting of the FL AVS and Florida Society for Microscopy.

Our goal is to get 50 people to join the FIB FIG within the first year. If
you are interested in becoming a member of the FIB FIG and/or interested in
attending the 1st FIB FIG Workshop in March, 2003 please contact either:

Lucille Giannuzzi, Chair, FIB FIG
lag-at-mail.ucf.edu

Joe Michael, Vice Chair, FIB FIG
jrmicha-at-sandia.gov

or

Barry Carter, Secretary/Treas, FIB FIG
carter-at-cems.umn.edu

*******************************************************************
Lucille A. Giannuzzi, Ph.D.

Associate Professor, Mechanical, Materials, and Aerospace Engineering
Director, UCF/Cirent Materials Characterization Facility
University of Central Florida
12443 Research Parkway, Suite 305
Orlando, FL 32826 email lag-at-mail.ucf.edu phone (407) 882-1500
fax (407) 275-4321
--------------------------------------------------------------------
"Good judgement comes from experience.

Experience comes from making bad judgement."

Mark Twain
********************************************************************




From daemon Fri Oct 11 10:10:20 2002



From: Marek Malecki, M.D., Ph.D., Professor :      mmalecki-at-facstaff.wisc.edu
Date: Fri, 11 Oct 2002 10:03:19 -0700
Subject: Re: polishing 1/2 " Cambridge carbon SEM mounts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Th. Bill:
What was ~ temp. which you were using?
Is that stuff conductive or you have to re-coat it?
Marek.


At 07:37 AM 10/11/02 -0400, you wrote:
} Apiezon W - a high temp vacuum wax, can be melted on to the surface and
} it will be perfectly smooth.
}
} Bill Miller
}
}
}
} At 06:20 PM 10/10/2002 -0700, you wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Fri Oct 11 10:10:20 2002



From: Marek Malecki, M.D., Ph.D., Professor :      mmalecki-at-facstaff.wisc.edu
Date: Fri, 11 Oct 2002 10:07:47 -0700
Subject: RE: polishing 1/2 " Cambridge carbon SEM mounts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Th. Wharton:
Please let me know after you try polishing it. We have all possible mills
and grinders here, but the pins , which we have, are really tough.
Marek.


At 07:22 AM 10/11/02 -0500, you wrote:

} Marek,
}
} Is this graphitic carbon? If so, you might look into vitreous carbon. This
} stuff is really hard, and shiny like a silicon wafer. I was looking for a
} really featureless, low Z mount, but not toxic (beryllium), and stumbled on
} this. I've just started using it so I haven't tried polishing it yet but I
} think it should be considerably easier to polish than graphitic carbon.
}
} Here's where I got it (sorry, not an EM supplier so you won't get standard
} mount geometries here).
}
} www.atomergic.com
}
} Regards,
} Wharton
}
} *************************************************************
} Wharton Sinkler, PhD
} UOP LLC
} 25 E. Algonquin Rd.
} Des Plaines, IL 6017-5017
} 847-391-3878
}
}
} } -----Original Message-----
} } From: Marek Malecki, M.D., Ph.D., Professor
} } [SMTP:mmalecki-at-facstaff.wisc.edu]
} } Sent: Thursday, October 10, 2002 8:21 PM
} } To: microscopy-at-sparc5.microscopy.com
} } Subject: polishing 1/2 " Cambridge carbon SEM mounts
} }
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Does anyone have attempted to polish /seal / ? carbon 1/2 " carbon SEM
} } mounts to obtain very smooth surface for HR SEM and be willing to share
} } the
} } procedure? Ultrafine-grain sand paper , carbon colloid, and card-board did
} }
} } not generate a surface which is smooth enough.



From daemon Fri Oct 11 14:23:45 2002



From: Gang Ning :      gning-at-mcw.edu
Date: Fri, 11 Oct 2002 14:11:03 -0500
Subject: Re: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi

I agree with Dr. Sara Miller's point that UA en block staining in some extent
helps in preservation of ultrastructure as well as enhancement of the
contrast. However, it also has disadvantages such as one has to deal with UA,
a biohazard reagent, and sometimes background from crystal precipitation,
especially when ethanolic UA is used. More importantly, it is time-consuming
because it takes hours for aqueous UA solution to penetrate a sizable tissue
block. Therefore, I would say it all depends your tissue and purpose of your
analysis.

In my lab, we routinely do en block with samples of blood, tissue culture,
virus, bacteria and other microorganisms, which are believed more vulnerable
to solvent, but not with solid tissue. We have a lot of samples from kidney
biopsy and we didn't find en block made any difference from staining thin
sections whereas taking your time to have a sufficient osmication is much
more helpful to your structure.

Greg Ning



mohammed y abdulrawoof / f40z006 75760 wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Dear listers,
} Please advise as to the usefulness/advantage of doing enblock staining of
} biological tissues as against doing heavy-metal staining of grids with
} thin sections. The tissues that I am about to process are for diagnosis
} (pathology).
}
} Thanks
} M. Yousuf Abdul-Rawoof

--
Gang Ning, M.D., Ph.D.
Director
Electron Microscopy Facility
Medical College of Wisconsin
8701 Watertown Plank Road
Milwaukee, Wisconsin 53226
(414) 456-8344
e-mail: gning-at-mcw.edu
http://www.mcw.edu/cellbio/emfacility.html




From daemon Fri Oct 11 15:24:45 2002



From: Ralph Common :      Ralph.Common-at-ht.msu.edu
Date: Fri, 11 Oct 2002 15:56:32 -0400
Subject: LM: Wild M400 Photomakroskop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I have recently acquired a Wild M400 Photomakroskop equipped for Polaroid
photography only. I would like to set it up for 35mm photography. I have
obtained an appropriate camera and controller, but the necessary relay lens
assembly (Wild part # 376 734) is no longer available. So far, I have been
unable to find anybody willing to part with one. Has anybody out there had
any success improvising a 35mm photography setup on the M400 using other
than the Wild original equipment? Any advice would be appreciated.

Ralph Common
Michigan State University
Division of Human Pathology
ralph.common-at-ht.msu.edu



From daemon Fri Oct 11 16:15:28 2002



From: Larry Stoter :      larry-at-cymru.freewire.co.uk
Date: Fri, 11 Oct 2002 20:57:27 +0100
Subject: RE: DPR

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


} Dear Mr. Udah;
}
} I will refer your request to a high powered government official in the
} Federal Prosecutor's Office in Newark, New Jersey.
}
} Sincerely,
}
} P. Tomic
}
} -----Original Message-----
} } From: ENGR. JOE UDAH [mailto:joeudah123-at-dpr-nig.com]
} Sent: Wednesday, October 09, 2002 11:49 AM
} To: microscopy-at-sparc5.microscopy.com
} Subject: DPR
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

snips fascinating but irrelevant

} further information on the procedures. Feel free to send response by Fax or
} TEL; expecting your response urgently. All enquiries should be directed to
} the undersigned by FAX OR PHONE.
} Looking forward to a good business relationship with you.
}
} Sincerely,
} ENGR. JOE UDAH.

Too late - the Enron management have already got it all stitched up :-))
--
Larry Stoter


From daemon Fri Oct 11 16:36:02 2002



From: William Oxberry :      William.Oxberry-at-downstate.edu (by way of
Date: Fri, 11 Oct 2002 16:29:43 -0500
Subject: Leica CPC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Listers-

We are considering purchasing the latest cryoworkstation (Leica CPC)
to plunge freeze samples into liquid propane in preparation for
freeze substitution. On our quote there was an automatic filling
device listed as optional and we decided not to purchase it to save
money (over $6000).
Now we are told that it is not optional but required because this
unit is not able to be filled manually.

Do any of you use this equipment and know if it can be filled
manually or needs an automatic filling device?

thanks in advance

Bill Oxberry
Downstate Medical Center
7182704472


From daemon Fri Oct 11 16:45:16 2002



From: saram-at-duke.edu
Date: Fri, 11 Oct 2002 17:35:39 -0400 (EDT)
Subject: Re: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


See CAPS interspersed below for distinction from the original.

Sara Miller

On Fri, 11 Oct 2002, Gang Ning wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi
}
} I agree with Dr. Sara Miller's point that UA en block staining in some extent
} helps in preservation of ultrastructure as well as enhancement of the
} contrast. However, it also has disadvantages such as one has to deal with UA,
} a biohazard reagent,

A BIOHAZARD IS A BIOLOGICAL AGENT THAT IS HAZARDOUS (INFECTIOUS). UA IS
RADIOACTIVE AND IS A HEAVY METAL. IT, LIKE GLUTARALDEHYDE, OSMIUM, LEAD,
RESINS, PROPYLENE OXIDE, ETC., SHOULD BE HANDLED WITH CARE. HANDLED
PROPERLY, IT IS NOT ANY MORE DANGEROUS THAN THE OTHER NASTY CHEMICALS
ELECTRON MICROSCOPISTS HAVE TO USE. IT IS NOT A BIOLOGICAL AGENT.

and sometimes background from crystal precipitation,
} especially when ethanolic UA is used.

WE NEVER SEE PRECIPITATION. PERHAPS TRY DILUTING THE STAIN OR WASHING
THOROUGHLY BEFORE USING IT WILL ELIMINATE THE PRECIPITATE. UA IS NOT
SOLUBLE IN PHOSPHATE OR CACODYLATE BUFFER; IF YOU DON'T WASH OUT THESE
BUFFERS BEFORE ADDING UA IN VERANAL ACETATE BUFFER, UA IN WATER, OR UA IN
ETHANOL OR METHANOL, YOU CAN GET PRECIPATE.

More importantly, it is time-consuming
} because it takes hours for aqueous UA solution to penetrate a sizable tissue
} block.

WE USE 1 MM TISSUE BLOCKS AND STAIN FOR AN HOUR 1% UA IN 0.11M VERONAL
ACETATE BUFFER. IF THE TISSUE OR CELL PELLET IS SMALLER, YOU CAN GET AWAY
WITH 30 MIN IN UA.

Therefore, I would say it all depends your tissue and purpose of your
} analysis.
}
I AGREE. IT ALL DEPENDS ON YOUR PURPOSE. IF YOU WANT BEAUTIFUL
ULTRASTRUCTURE AND YOU'RE NOT LOOKING SPECIFICALLY FOR GLYCOGEN, USE UA EN
BLOCK. IF YOU'RE IN A HURRY OR YOU CARE TO SEE ALL THE GLYGOGEN, DON'T
USE IT.

} In my lab, we routinely do en block with samples of blood, tissue culture,
} virus, bacteria and other microorganisms, which are believed more vulnerable
} to solvent, but not with solid tissue. We have a lot of samples from kidney
} biopsy and we didn't find en block made any difference from staining thin
} sections whereas taking your time to have a sufficient osmication is much
} more helpful to your structure.
}
} Greg Ning
}
}
}
} mohammed y abdulrawoof / f40z006 75760 wrote:
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} } Dear listers,
} } Please advise as to the usefulness/advantage of doing enblock staining of
} } biological tissues as against doing heavy-metal staining of grids with
} } thin sections. The tissues that I am about to process are for diagnosis
} } (pathology).
} }
} } Thanks
} } M. Yousuf Abdul-Rawoof
}
} --
} Gang Ning, M.D., Ph.D.
} Director
} Electron Microscopy Facility
} Medical College of Wisconsin
} 8701 Watertown Plank Road
} Milwaukee, Wisconsin 53226
} (414) 456-8344
} e-mail: gning-at-mcw.edu
} http://www.mcw.edu/cellbio/emfacility.html
}
}
}
}
}

Sara E. Miller, Ph. D.
P. O. Box 3712
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-3265



From daemon Fri Oct 11 20:48:00 2002



From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Fri, 11 Oct 2002 15:27:51 -1000 (HST)
Subject: Re: polishing 1/2 " Cambridge carbon SEM mounts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi, Marek-

} Does anyone have attempted to polish /seal / ? carbon 1/2 " carbon SEM
} mounts to obtain very smooth surface for HR SEM and be willing to share the
} procedure? Ultrafine-grain sand paper , carbon colloid, and card-board did
} not generate a surface which is smooth enough.

We have a user who gets excellent results by gluing a small piece of
aluminum foil onto the stub and burnishing it before mounting her
micromolluscs. The background is amazingly smooth. I've not tried to
duplicate her technique, so I can't claim it will work all the time and
with all brands of foil. I have seen machine marks on some foils. I'm also
not sure if she uses regular thickness or heavy-duty. I will email her for
details when she returns from the field (Tahiti)!

Bug me if you don't hear from me.

Aloha,
Tina


****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************



From daemon Sat Oct 12 09:50:09 2002



From: Marek Malecki, M.D., Ph.D., Professor :      mmalecki-at-facstaff.wisc.edu
Date: Sat, 12 Oct 2002 09:45:51 -0700
Subject: Re: polishing 1/2 " Cambridge carbon SEM mounts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Thanks Tina:
Looking forward to the details.
We were attempting to polish Al stubs, but they came out not flat enough.
Marek.

PS
How is your 912 working?



At 03:27 PM 10/11/02 -1000, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Sat Oct 12 11:03:56 2002



From: Marek Malecki, M.D., Ph.D., Professor :      mmalecki-at-facstaff.wisc.edu
Date: Sat, 12 Oct 2002 11:00:09 -0700
Subject: HR EDX elemental mapping programs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello EDX Experts;
Would you be willing to share your experience with various EDX mapping
programs - in particular deconvolving elemental distribution with
corrections for binding and emission energies in alloys or particles
embedded in a matrix (C)?
Marek.


From daemon Sat Oct 12 13:54:23 2002



From: ggm-at-servidor.unam.mx (by way of Ask-A-Microscopist)
Date: Sat, 12 Oct 2002 13:48:23 -0500
Subject: Ask-A-Microscopist:SEM process de bacteria Tiobacillus

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (ggm-at-servidor.unam.mx) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday,
October 11, 2002 at 11:39:52
---------------------------------------------------------------------------

Email: ggm-at-servidor.unam.mx
Name: GUILLERMINA GLEZ.

Organization: UNAM

Education: Graduate College

Location: MEXICO, D.F., MEXICO.

Question: do you now how to process de bacteria Tiobacillus sulfurans to SEM?

especifilly, how to concentrate, and conditions to concentrate.

thanks.

---------------------------------------------------------------------------


From daemon Sat Oct 12 13:54:26 2002



From: gbarclay-at-trinidad.net (by way of Ask-A-Microscopist)
Date: Sat, 12 Oct 2002 13:47:53 -0500
Subject: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (gbarclay-at-trinidad.net) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday,
October 11, 2002 at 23:40:22
---------------------------------------------------------------------------

Email: gbarclay-at-trinidad.net
Name: Greg Barclay

Organization: University of the West Indies

Education: Graduate College

Location: St. Augustine, Trinidad, West Indies

Question: Light microscope terminology: We all know what a
stereomicroscope is and what a compound microscope is, but are they
not technically both the same thing? Should not the distinction be
between the "simple microscope," basically a highly corrected doublet
or triplet lens, forming the Dowland type of microscope, and any kind
of microscope with a separate eyepiece and objective lens? If so,
then both the stereo and binocular (or monocular) microscopes are
compond microscopes. I have been referring to "stereomicroscopes and
binocular microscopes" in our Life Sciences Dept, and having some
arguments with those who refer "stereomicroscopes and compound
microscopes," over what is really what.

---------------------------------------------------------------------------


From daemon Sat Oct 12 16:09:14 2002



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sat, 12 Oct 2002 16:57:48 -0500
Subject: Smooth carbon surfaces

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Marek Malecki wrote:
================================================================
Does anyone have attempted to polish /seal / ? carbon 1/2 " carbon SEM
mounts to obtain very smooth surface for HR SEM and be willing to share the
procedure? Ultrafine-grain sand paper , carbon colloid, and card-board did
not generate a surface which is smooth enough.
================================================================
Highly Ordered Pyrolytic Graphite (HOPG) which can be freshly cleaved like
mica, will cleave to an atomically smooth surface, mirror reflective. The
size of the atomically smooth areas depend on the grade of HOPG that is
procured. While the cost at first might seem high, remember one can get a
number of cleavings out of one HOPG starting "block". This is the substrate
often times used by those doing SPM work because of its ultra smooth surface
relative to other ways to make smooth nonpolar surfaces in carbon or
graphite.

In addition, it is of high purity and other wise behaves like any other form
of carbon in terms of its chemical resistance and stability.

More information about HOPG can be found on URL
http://www.2spi.com/catalog/new/hopgsub.shtml

Disclaimer: SPI Supplies is a leading supplier of HOPG to microscopy
laboratories worldwide so we have a vested interest in seeing more of it
used in this kind of application.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================




From daemon Sat Oct 12 17:37:46 2002



From: qualityimages :      qualityimages-at-netrax.net
Date: Sat, 12 Oct 2002 18:25:17 -0400
Subject: Re: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


It seems to me that there jas always been one other difference between
the two. Most truly binocular (stereo) microscopes do not have an
inverted image, whereas most "compound" microscopes have an inverted image.

Ken Converse
owner
Quality Images
third party SEM service
Delta, PA

by way of Ask-A-Microscopist wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (gbarclay-at-trinidad.net) from
} http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday,
} October 11, 2002 at 23:40:22
} ---------------------------------------------------------------------------
}
}
} Email: gbarclay-at-trinidad.net
} Name: Greg Barclay
}
} Organization: University of the West Indies
}
} Education: Graduate College
}
} Location: St. Augustine, Trinidad, West Indies
}
} Question: Light microscope terminology: We all know what a
} stereomicroscope is and what a compound microscope is, but are they
} not technically both the same thing? Should not the distinction be
} between the "simple microscope," basically a highly corrected doublet
} or triplet lens, forming the Dowland type of microscope, and any kind
} of microscope with a separate eyepiece and objective lens? If so, then
} both the stereo and binocular (or monocular) microscopes are compond
} microscopes. I have been referring to "stereomicroscopes and binocular
} microscopes" in our Life Sciences Dept, and having some arguments with
} those who refer "stereomicroscopes and compound microscopes," over
} what is really what.
}
} ---------------------------------------------------------------------------
}
}
}





From daemon Sun Oct 13 00:54:20 2002



From: hgccgfdcgf6252b45-at-libero.it
Date: Sat, 12 Oct 0102 12:11:15 +0800
Subject: Viag and other products 7468w-5

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html
charset="iso-8859-1"
Content-Transfer-Encoding: 8bit
X-Priority: 3 (Normal)
X-MSMail-Priority: Normal
X-Mailer: Microsoft Outlook, Build 10.0.2616
Importance: Normal



From: hgccgfdcgf6252b45-at-libero.it
Date: Sat, 12 Oct 0102 12:11:15 +0800
Subject: Viag and other products 7468w-5

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


We Offer The Widest Range Of Prescription Drugs

Available Through Online Ordering


http://www.great-rx.com/main2.php?rx=18080


Phentermine (Weight-Loss)

Celebrex (Pain-Relief)

Valtrex (Treatment for Herpes)

Meridia (Weight-Loss)

Xenical (Weight-Loss

Viagra (Sexual)

Phentermine (Weight-loss)

Zyban (Stop Smoking)

Plus Many Other Prescription Drugs!



http://www.great-rx.com/main2.php?rx=18080


Enjoy complete privacy, discretion, and dignity while
addressing your healthcare needs.


All packages shipped in plain packaging to protect
your privacy!


http://www.great-rx.com/main2.php?rx=18080













to get off this list.
http://greenzer.com/remove.php
1086OneA4-219nCYh9916DCJm7-930Bl29


From daemon Sun Oct 13 15:34:22 2002



From: saram-at-duke.edu
Date: Sun, 13 Oct 2002 16:15:09 -0400 (EDT)
Subject: Re: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) id PAA24924
for dist-Microscopy; Sun, 13 Oct 2002 15:21:56 -0500 (CDT)
Received: from njz_spm_filter (sparc5 [206.69.208.10])
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with SMTP id PAA24920
for "MicroscopyFilteredEmail2-at-msa.microscopy.com"; Sun, 13 Oct 2002 15:21:26 -0500 (CDT)
by sparc5.microscopy.com (8.9.3+Sun/8.9.3) with ESMTP id PAA24913
for {Microscopy-at-sparc5.microscopy.com} ; Sun, 13 Oct 2002 15:21:14 -0500 (CDT)
Received: from godzilla6.acpub.duke.edu (godzilla6.acpub.duke.edu [152.3.102.8])
by gibson.acpub.duke.edu (8.11.5/8.11.3/Duke-5.0.0) with ESMTP id g9DKFAO11358;
Sun, 13 Oct 2002 16:15:10 -0400 (EDT)
Received: (from saram-at-localhost)
by godzilla6.acpub.duke.edu (8.9.3/8.9.3) id QAA02772;
Sun, 13 Oct 2002 16:15:09 -0400 (EDT)


Again, see CAPS below.
(CAPS ARE USED TO DIFFERENTIATE COMMENTS, NOT TO CONVEY HOSTILITY.)


On Sun, 13 Oct 2002, gang ning wrote:

} Thanks for the clear definition of biohazard. But it is obvious that
} 1) UA is a nasty/hazard chemical. Everyone wants to avoid it if possible,

UA CERTAINLY CAN BE CALLED NASTY. IT DOES REQUIRE SPECIAL HANDLING
PROCEDURES, AS DO OTHER CHEMICALS THAT WE USE AND CAN BE HANDLED SAFELY
UNDER THE PROPER CONDITIONS. NOT EVERYONE WANTS TO AVOID IT. FOLKS MUST
MAKE THEIR OWN JUDGEMENT AS TO WHETHER THE EXTRA PROCESSING TIME AND CARE
OF HANDLING ANOTHER CHEMICAL IS WORTH IT TO THEM. NOT USING UA IS FINE.
HOWEVER, DO NOT MAKE A GENERALIZATION ABOUT *EVERYONE* JUST BECAUSE YOU
CHOOSE NOT TO USE IT.

} 2) it does take at least 1 hr + time for washing thoroughly (how long?)
} for a 1 mm3 block .... it is clearly time-consuming procedure

IT DOES REQUIRE EXTRA TIME (1-2 HOURS). ONE CAN STAIN A 1 mm3 BLOCK FOR
30-60 MIN AND WASH FOR 30-60 MIN.

unless your turn around time is not an issue and working overtime is not a
} matter with your hospital policy,

TIME IS VERY MUCH AN ISSUE FOR US IN THE HOSPITAL SETTING. OUR TURN
AROUND TIME IS *ROUTINELY* NEXT DAY FOR SAMPLES FIXED IN GLUT, OS, UA.
TISSUE COMES IN IN THE AM, IS PROCESSED THAT DAY, BAKED OVERNIGHT, CUT AND
VIEWED THE NEXT DAY, AND PATHOLOGISTS HAVE MICROGRAPHS *IN HAND* THAT
AFTERNOON--ALMOST ALWAYS. MY TECHS RARELY HAVE TO WORK OVERTIME, AND WHEN
THEY DO, IT'S USUALLY BECAUSE OF EXTRA CUTTING OF HAVING TO RETRIEVE
SPECIMENS OUT OF PARAFFIN BLOCKS OR OFF SLIDES, NOT BECAUSE THEY PUT
SOMETHING INTO UA.

YOU ARE CLEARLY VERY NEGATIVE TOWARD THE USE OF UA. FINE. DON'T USE IT.
BUT DON'T BE SARCASTIC ABOUT SPECIMEN TURN AROUND TIME AND
TECHNICIAN OVERTIME, BECAUSE IT JUST ISN'T AN ISSUE IF SPECIMENS AND
WORKLOADS ARE MANAGED PROPERLY.

} 3) en block staining has adversary effect on showing glycogen so is not
} suitable for muscle, heart, liver and so on; there is not significant
} meaning for kidney .. then what else left? LOTS:

BRAIN, CILIA, TUMORS, VIRUSES, MUSCLE DISEASES OTHER THAN GLYCOGEN STORAGE
DISEASES, RESEARCH SAMPLES, ETC.


It is overkill for solid tissue blocks, IT IS NOT OVERKILL WHEN THE
BEST PRESERVATION, RATHER THAN DECREASING PROCESSING TIME IS OF
IMPORTANCE.

and 4) ethanolic/methanolic UA of higher concentration penetrates tissue
} much faster than other aqueous UA solutions and gives fresher image and
} better contrast but sometimes leave precipitation around cell member no
} matter how one washes before staining.

AS I SAID BEFORE, IF THE CORRECT BUFFER IS USED SO THAT UA DOES NOT
PRECIPTATE WITH IT, THERE IS NO PRECIPITATION. PERHAPS YOUR *HIGHER
CONCENTRATION* IS A CAUSE OF DIFFICULTY IN WASHING. SINCE UA IS SOLUBLE
IN ALCOHOL, UNBOUND, UNPRECIPITATED UA IS EASILY WASHED OUT BY THE
DEHYDRATING AGENT(S). IF YOU HAVE PRECIPITATE, IT IS MOST LIKELY DUE TO
SOME REACTION WITH THE SOLUTION LEFT IN THE CELL BEFORE YOU ADD THE UA, OR
PERHAPS THE *HIGHER* CONCENTRATION YOU USE ISN'T WASHED OUT???

} For EM work, the simpler and better, as long as it works.

YES, THE KEY IS, *AS LONG AS IT WORKS* FOR THE PURPOSE YOU HAVE. IF THE
PURPOSE IS TO PRODUCE THE BEST ULTRASTRUCTURE, AND IF MEMBRANES AND OR
LIPIDS ARE PART OF THE QUESTION, USE OF UA IS GOOD.

Overkill is always a problem. YES, NO ONE WANTS TO SPEND MORE TIME THAN
IS NECESSARY OR USE CHEMICALS THAT CAN BE HAZARDOUS IF NOT HANDLED
PROPERLY, BUT UA IS NOT ALWAYS OVERKILL.

IT DEPENDS ON YOUR APPLICATION. EACH TO HIS OWN.
}
}
}
} saram-at-duke.edu {mailto:saram-at-duke.edu} wrote:
}
} } See CAPS interspersed below for distinction from the original.
} }
} } Sara Miller
} }
} } On Fri, 11 Oct 2002, Gang Ning wrote:
} }
} } } ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com {mailto:ListServer-at-MSA.Microscopy.Com}
} } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------.
} } }
} } }
} } } Hi
} } }
} } } I agree with Dr. Sara Miller's point that UA en block staining in some extent
} } } helps in preservation of ultrastructure as well as enhancement of the
} } } contrast. However, it also has disadvantages such as one has to deal with UA,
} } } a biohazard reagent,
} } }
} }
} } A BIOHAZARD IS A BIOLOGICAL AGENT THAT IS HAZARDOUS (INFECTIOUS). UA IS
} } RADIOACTIVE AND IS A HEAVY METAL. IT, LIKE GLUTARALDEHYDE, OSMIUM, LEAD,
} } RESINS, PROPYLENE OXIDE, ETC., SHOULD BE HANDLED WITH CARE. HANDLED
} } PROPERLY, IT IS NOT ANY MORE DANGEROUS THAN THE OTHER NASTY CHEMICALS
} } ELECTRON MICROSCOPISTS HAVE TO USE. IT IS NOT A BIOLOGICAL AGENT.
} }
} } and sometimes background from crystal precipitation,
} }
} } } especially when ethanolic UA is used.
} } }
} }
} } WE NEVER SEE PRECIPITATION. PERHAPS TRY DILUTING THE STAIN OR WASHING
} } THOROUGHLY BEFORE USING IT WILL ELIMINATE THE PRECIPITATE. UA IS NOT
} } SOLUBLE IN PHOSPHATE OR CACODYLATE BUFFER; IF YOU DON'T WASH OUT THESE
} } BUFFERS BEFORE ADDING UA IN VERANAL ACETATE BUFFER, UA IN WATER, OR UA IN
} } ETHANOL OR METHANOL, YOU CAN GET PRECIPATE.
} }
} } More importantly, it is time-consuming
} }
} } } because it takes hours for aqueous UA solution to penetrate a sizable tissue
} } } block.
} } }
} }
} } WE USE 1 MM TISSUE BLOCKS AND STAIN FOR AN HOUR 1% UA IN 0.11M VERONAL
} } ACETATE BUFFER. IF THE TISSUE OR CELL PELLET IS SMALLER, YOU CAN GET AWAY
} } WITH 30 MIN IN UA.
} }
} } Therefore, I would say it all depends your tissue and purpose of your
} }
} } } analysis.
} } }
} } I AGREE. IT ALL DEPENDS ON YOUR PURPOSE. IF YOU WANT BEAUTIFUL
} } ULTRASTRUCTURE AND YOU'RE NOT LOOKING SPECIFICALLY FOR GLYCOGEN, USE UA EN
} } BLOCK. IF YOU'RE IN A HURRY OR YOU CARE TO SEE ALL THE GLYGOGEN, DON'T
} } USE IT.
} }
} } } In my lab, we routinely do en block with samples of blood, tissue culture,
} } } virus, bacteria and other microorganisms, which are believed more vulnerable
} } } to solvent, but not with solid tissue. We have a lot of samples from kidney
} } } biopsy and we didn't find en block made any difference from staining thin
} } } sections whereas taking your time to have a sufficient osmication is much
} } } more helpful to your structure.
} } }
} } } Greg Ning
} } }
} } }
} } }
} } } mohammed y abdulrawoof / f40z006 75760 wrote:
} } }
} } } } ------------------------------------------------------------------------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com {mailto:ListServer-at-MSA.Microscopy.Com}
} } } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } -----------------------------------------------------------------------.
} } } }
} } } } Dear listers,
} } } } Please advise as to the usefulness/advantage of doing enblock staining of
} } } } biological tissues as against doing heavy-metal staining of grids with
} } } } thin sections. The tissues that I am about to process are for diagnosis
} } } } (pathology).
} } } }
} } } } Thanks
} } } } M. Yousuf Abdul-Rawoof
} } } }
} } } --
} } } Gang Ning, M.D., Ph.D.
} } } Director
} } } Electron Microscopy Facility
} } } Medical College of Wisconsin
} } } 8701 Watertown Plank Road
} } } Milwaukee, Wisconsin 53226
} } } (414) 456-8344
} } } e-mail: gning-at-mcw.edu {mailto:gning-at-mcw.edu}
} } } http://www.mcw.edu/cellbio/emfacility.html
} } }
} } }
} } }
} } }
} } }
} }
} } Sara E. Miller, Ph. D.
} } P. O. Box 3712
} } Duke University Medical Center
} } Durham, NC 27710
} } Ph: 919 684-3452
} } FAX: 919 684-3265
} }
} }
}
} --
}
} Gang Ning, M.D., Ph.D.
}
} Director
}
} Electron Microscopy Facility
}
} Medical College of Wisconsin
}
} 8701 Watertown Plank Road
}
} Milwaukee, WI 53226
}
} (414) 456-8141
}
}
}
}

Sara E. Miller, Ph. D.
P. O. Box 3712
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-3265



From daemon Sun Oct 13 17:47:31 2002



From: Rosemary White :      rosemary.white-at-csiro.au
Date: Mon, 14 Oct 2002 08:30:58 +1000
Subject: Re: polishing 1/2 " Cambridge carbon SEM mounts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Another option is to glue a small round glass coverslip onto the stub with
carbon paste, then coat with gold, then attach specimens, re-coat.

} Hi, Marek-
}
} } Does anyone have attempted to polish /seal / ? carbon 1/2 " carbon SEM
} } mounts to obtain very smooth surface for HR SEM and be willing to share the
} } procedure? Ultrafine-grain sand paper , carbon colloid, and card-board did
} } not generate a surface which is smooth enough.
}
} We have a user who gets excellent results by gluing a small piece of
} aluminum foil onto the stub and burnishing it before mounting her
} micromolluscs. The background is amazingly smooth. I've not tried to
} duplicate her technique, so I can't claim it will work all the time and
} with all brands of foil. I have seen machine marks on some foils. I'm also
} not sure if she uses regular thickness or heavy-duty. I will email her for
} details when she returns from the field (Tahiti)!
}
} Bug me if you don't hear from me.
}
} Aloha,
} Tina
}
}
} ****************************************************************************
} * Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
} * Biological Electron Microscope Facility * (808) 956-6251 *
} * University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
} ****************************************************************************




From daemon Sun Oct 13 19:56:08 2002



From: John V Nailon :      J.Nailon-at-mailbox.uq.edu.au
Date: Mon, 14 Oct 2002 10:48:52 +1000
Subject: Re: polishing 1/2 " Cambridge carbon SEM mounts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


We have a preference for mounting a coated 0.5" coverslip onto the stub
with colloidal dag. The cover slip is coated with poly-L-lysine either
before or after mounting on the stub. This holds the sample well and
gives a smooth black background after sputter coating.
Regards
JVN

Tina Carvalho wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi, Marek-
}
}
} } Does anyone have attempted to polish /seal / ? carbon 1/2 " carbon SEM
} } mounts to obtain very smooth surface for HR SEM and be willing to share the
} } procedure? Ultrafine-grain sand paper , carbon colloid, and card-board did
} } not generate a surface which is smooth enough.
} }
}
} We have a user who gets excellent results by gluing a small piece of
} aluminum foil onto the stub and burnishing it before mounting her
} micromolluscs. The background is amazingly smooth. I've not tried to
} duplicate her technique, so I can't claim it will work all the time and
} with all brands of foil. I have seen machine marks on some foils. I'm also
} not sure if she uses regular thickness or heavy-duty. I will email her for
} details when she returns from the field (Tahiti)!
}
} Bug me if you don't hear from me.
}
} Aloha,
} Tina
}
}
} ****************************************************************************
} * Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
} * Biological Electron Microscope Facility * (808) 956-6251 *
} * University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
} ****************************************************************************
}
}
}
}
}


--
John V Nailon
Executive Officer and Operations Manager
The Centre for Microscopy and Microanlaysis
The University of Queensland
St Lucia QLD 4072
Tel: +61-7-33654214
Fax: +61-7-33654422
WWW: http://www.uq.edu.au/nanoworld



From daemon Mon Oct 14 06:27:56 2002



From: daniel eberhard :      daniel.eberhard-at-uni-bielefeld.de
Date: Mon, 14 Oct 2002 13:16:55 +0200
Subject: smooth muscle anti actin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,
I´m searching for a smooth muscle anti-Actin antibody (recognizing mouse
smooth muscle) which also works on formaldehyde fixed tissue sections.
Any hints or recommendations are welcome.

Thanks in advance
Daniel



From daemon Mon Oct 14 07:42:35 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Mon, 14 Oct 2002 08:41:01 -0400
Subject: Re: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I must jump in on this one. A real "stereo" microscope has two parallel
optical paths. That is, two objective lenses. No microscope with one
objective lens can be "stereo." Albeit there may be two eyepieces but
without two separate optical paths the images in both eyepieces must
necessarily be identical. It's like listening to stereo sound but if the
left and right channels are identical, hence no depth or direction to the
sound.

Inverted or non-inverted images do not define a compound microscope since
many are made with an "image erector" which is simply an additional lens to
reverse the image so you view the sample as it is with the unaided eye.
Most "micromanipulator" systems have image erectors on them for that
purpose. It helps one to not have to re-wire one's neurons to interpret
left as right when having to use hand manipulation under the microscope. I
have found that wiring and re-wiring neurons problematic.

Peter
Anadigics, Inc.

-----Original Message-----
} From: qualityimages [mailto:qualityimages-at-netrax.net]
Sent: Saturday, October 12, 2002 6:25 PM
To: by way of Ask-A-Microscopist; Microscopy


It seems to me that there jas always been one other difference between
the two. Most truly binocular (stereo) microscopes do not have an
inverted image, whereas most "compound" microscopes have an inverted image.

Ken Converse
owner
Quality Images
third party SEM service
Delta, PA

by way of Ask-A-Microscopist wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (gbarclay-at-trinidad.net) from
} http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday,
} October 11, 2002 at 23:40:22
}
---------------------------------------------------------------------------
}
}
} Email: gbarclay-at-trinidad.net
} Name: Greg Barclay
}
} Organization: University of the West Indies
}
} Education: Graduate College
}
} Location: St. Augustine, Trinidad, West Indies
}
} Question: Light microscope terminology: We all know what a
} stereomicroscope is and what a compound microscope is, but are they
} not technically both the same thing? Should not the distinction be
} between the "simple microscope," basically a highly corrected doublet
} or triplet lens, forming the Dowland type of microscope, and any kind
} of microscope with a separate eyepiece and objective lens? If so, then
} both the stereo and binocular (or monocular) microscopes are compond
} microscopes. I have been referring to "stereomicroscopes and binocular
} microscopes" in our Life Sciences Dept, and having some arguments with
} those who refer "stereomicroscopes and compound microscopes," over
} what is really what.
}
}
---------------------------------------------------------------------------
}
}
}





From daemon Mon Oct 14 08:08:49 2002



From: Barbara Foster :      bfoster-at-mme1.com (by way of MicroscopyListserver)
Date: Mon, 14 Oct 2002 08:03:55 -0500
Subject: Re: EM job classification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



}
} }
} } Hi,
} }
} } I usually don't weigh in on this type of thing but, as a past negotiator
} } for the teachers in my old school system and as an ardent advocate of the
} } importance of experience in microscopy, here is my two cents:
} } 1. As a negotiator - I worked very closely with both the Mass Teachers
} } Association (MTA) and the National Teachers Association. Ironically, our
} } "supervisors" (curriculum supervisors, principals and vice principals)
} } needed the protection of our association (MTA and NEA preferred not to be
} } called unions but they filled the role of protectors and negotiators), so,
} } to offer them that support, we opened membership to them but, for
} } negotiation purposes, separated them into their own negotiating
} group. The
} } position of lab director
} }
} } 2. As for your 20 years' experience: I earn my living training
} } microscopists. I head a group of consultants who specialize in
} } training. I am also a contributing editor to two key publications. Our
} } company is THE primary source of market research in the microscopy and
} } imaging industry. With well over 20 year's experience in the field as a
} } microscopist, a trainer, and now, a recognized industry watcher, I will
} } tell you the following:
} } The general rule of thumb is that it takes, on average, at least a year
} for
} } a microscopist to become proficient. Why?
} } a.Unlike any other analytical discipline, microscopy demands a high level
} } of eye-brain-hand coordination.
} } b. Unlike most other disciplines, we don't have one or two major pieces of
} } instrumentation to master; we have multiple instrument systems. We
} have to
} } know how to prepare our samples, which type of microscopy to use, how to
} } interpret the myriad types of images each type of microscopy can produce,
} } and, finally, how to integrate a complex electronic system including
} } cameras, computers, and software, to collect, archive, process, and
} measure
} } our images.
} } c. Unlike most other disciplines, there is no replacement for the mental
} } library of images which a microscopist collects with experience. Yes, a
} } spectroscopist will learn the nuances of an upfield or downfield shift
} } characteristic of a specific spectrum, but no other discipline requires
} the
} } mental collection and correlation of image information like microscopy.
} }
} } As a master teacher and contributing editor, I know that I must weigh my
} } words carefully. However, as you can tell, I feel very strongly about the
} } erosion of microscopy experience which inevitably accompanies budget
} } cuts. Please feel free to quote me to your management on any or all of
} } these comments.
} }
} } Best regards... and strong support....
} } Barbara Foster
} } Microscopy/Microscopy Education
} } 125 Paridon Street, Suite 102
} } Springfield, MA 01118
} } PH: 413-746-6931 FX: 413-746-9311 Web: www.MicroscopyEducation.com
} }
} }
} }
} }
} } At 06:43 PM 10/10/02 -0400, Ed Calomeni wrote:
} } } ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------.
} } }
} } }
} } } Hi Dorota,
} } }
} } } Fight this with all you have.
} } }
} } } The institution I just quit downgraded all of the lab staff positions. A
} } } microbiologist with no em experience replaced me with 20 years of em
} } } experience only because she lost her job due to downsizing and bumped
} me out
} } } of my position. The management had recently changed (lowered) all of the
} } } standards/qualifications so this event could take place. The union was
} } } never informed of this move, but that is another story.
} } }
} } } A supervisors position should not be part of the union. A smart manager
} } } should realize that you cannot hire good help in this field into a
} } } supervisor position without at least 7-10 years of EM experience.
} } }
} } } The minimum education should be at least a BS in a biological field
} with at
} } } least 5 years experience.
} } }
} } } Cheers,
} } }
} } } Ed Calomeni
} } }
} } }
} } }
} } } ----- Original Message -----
} } } } From: "Dorota Wadowska" {wadowska-at-upei.ca}
} } } To: {microscopy-at-sparc5.microscopy.com}
} } } Sent: Tuesday, October 08, 2002 10:09 AM
} } } Subject: EM job classificcation
} } }
} } }
} } } }
} ------------------------------------------------------------------------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } } } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } } } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } }
} -----------------------------------------------------------------------.
} } } }
} } } }
} } } } Hi Listservers!
} } } } I have a few questions that I would like to ask you If you feel you
} } } } know the answer please reply. My university together with local
} } } } union reclassified staff positions according to Aiken plan. Needless
} } } } to say my position was downgraded.
} } } } My questions are:
} } } } 1. Can EM supervisory position be compared to any on campus
} } } } staff positions (university has Department of Sciences and
} } } } Veterinary Medecine)? I supervise small multi--user lab, we have 2
} } } } TEM, and we do research, diagnostics, teaching (under- and
} } } } graduate level) in the field of biological sciences. No fancy staff
} like
} } } } X-ray analysis or cryo.
} } } } 2. What in your opinion is a necessary minimum experience
} } } } required to perform this job ? I am talking about accepting the
} } } } position and immediately getting into the stream of work.
} } } } 3. What in your opinion is the minimum level of education that is
} } } } required to perform this job?
} } } }
} } } } You can reply directly to me. Thanks in advance
} } } } Frustrated
} } } } Dorota
} } } }
} }

{/x-flowed}


From daemon Mon Oct 14 10:12:03 2002



From: John W. Raffensperger, Jr. :      johnr-at-helwigcp.com
Date: Mon, 14 Oct 2002 09:58:57 -0500
Subject: RE: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Just because a microscope has two eyepieces, does not mean that it is a
"stereo" microscope.

A stereo microscope has not only two eyepieces, but also two objectives,
and two separate light paths. Because of having two slightly offset
images, you see a three dimensional image. Usually these are lower
powered systems, but still compound microscopes (multiple lens systems).

On a high power (relatively) microscope with a binocular head, you only
have one objective, and one light path from the specimen to the head.
The prism(s) in the head divide the light path to provide multiple
images, but they are all (three in the case of an extra camera port)
identical. This does not provide a 3-D image, it just makes it easier
on the eyes and brain to process the image because one eye is not
distracted.

I am not a scientist, and would be interested in any corrections to the
above.

John W. Raffensperger, Jr.
IS Manager
Helwig Carbon Products, Inc.


Email: gbarclay-at-trinidad.net
Name: Greg Barclay

Organization: University of the West Indies

Education: Graduate College

Location: St. Augustine, Trinidad, West Indies

Question: Light microscope terminology: We all know what a
stereomicroscope is and what a compound microscope is, but are they
not technically both the same thing?



From daemon Mon Oct 14 11:31:52 2002



From: Isabel Nogueira :      isabeln-at-popsrv.ist.utl.pt
Date: Mon, 14 Oct 2002 17:16:01 +0100
Subject: Glue for TEM samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I need to prepare TEM cross-section samples of alternate SiO2 and TiO2
layers on a Si substrate. For that purpose I would appreciate suggestions on
glues, suitable for TEM work, that have a sputtering rate not too different
from my non homogeneous layers. The final sputtering will be on an Ion Mill
(4keV) using Ar+ ions.
All suggestions are welcome (including vendors) either to the list or to me
directly.
Thank you,

Isabel



Isabel Nogueira
Instituto Superior Técnico
Dep. Materiais
Avenida Rovisco Pais
1049-001 Lisboa
Portugal
tel.: +351 218418123
fax: +351 218418120
email: isabeln-at-popsrv.ist.utl.pt



From daemon Mon Oct 14 13:25:36 2002



From: Barbara Foster :      bfoster-at-mme1.com
Date: Mon, 14 Oct 2002 14:23:57 -0700
Subject: Fwd: Re: EM job classification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html




} From: Barbara Foster {bfoster-at-mme1.com}
} Subject: Re: EM job classification
}
} }
} } }
} } } Hi,
} } }
} } } I usually don't weigh in on this type of thing but, as a past negotiator
} } } for the teachers in my old school system and as an ardent advocate of the
} } } importance of experience in microscopy, here is my two cents:
} } } 1. As a negotiator - I worked very closely with both the Mass Teachers
} } } Association (MTA) and the National Teachers Association. Ironically, our
} } } "supervisors" (curriculum supervisors, principals and vice principals)
} } } needed the protection of our association (MTA and NEA preferred not to be
} } } called unions but they filled the role of protectors and negotiators),
} } so,
} } } to offer them that support, we opened membership to them but, for
} } } negotiation purposes, separated them into their own negotiating
} } group. The
} } } position of lab director
} } }
} } } 2. As for your 20 years' experience: I earn my living training
} } } microscopists. I head a group of consultants who specialize in
} } } training. I am also a contributing editor to two key publications. Our
} } } company is THE primary source of market research in the microscopy and
} } } imaging industry. With well over 20 year's experience in the field as a
} } } microscopist, a trainer, and now, a recognized industry watcher, I will
} } } tell you the following:
} } } The general rule of thumb is that it takes, on average, at least a
} } year for
} } } a microscopist to become proficient. Why?
} } } a.Unlike any other analytical discipline, microscopy demands a high level
} } } of eye-brain-hand coordination.
} } } b. Unlike most other disciplines, we don't have one or two major
} } pieces of
} } } instrumentation to master; we have multiple instrument systems. We
} } have to
} } } know how to prepare our samples, which type of microscopy to use, how to
} } } interpret the myriad types of images each type of microscopy can produce,
} } } and, finally, how to integrate a complex electronic system including
} } } cameras, computers, and software, to collect, archive, process, and
} } measure
} } } our images.
} } } c. Unlike most other disciplines, there is no replacement for the mental
} } } library of images which a microscopist collects with experience. Yes, a
} } } spectroscopist will learn the nuances of an upfield or downfield shift
} } } characteristic of a specific spectrum, but no other discipline
} } requires the
} } } mental collection and correlation of image information like microscopy.
} } }
} } } As a master teacher and contributing editor, I know that I must weigh my
} } } words carefully. However, as you can tell, I feel very strongly about
} } the
} } } erosion of microscopy experience which inevitably accompanies budget
} } } cuts. Please feel free to quote me to your management on any or all of
} } } these comments.
} } }
} } } Best regards... and strong support....
} } } Barbara Foster
} } } Microscopy/Microscopy Education
} } } 125 Paridon Street, Suite 102
} } } Springfield, MA 01118
} } } PH: 413-746-6931 FX: 413-746-9311 Web: www.MicroscopyEducation.com
} } }
} } }
} } }
} } }
} } } At 06:43 PM 10/10/02 -0400, Ed Calomeni wrote:
} } } } ------------------------------------------------------------------------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } -----------------------------------------------------------------------.
} } } }
} } } }
} } } } Hi Dorota,
} } } }
} } } } Fight this with all you have.
} } } }
} } } } The institution I just quit downgraded all of the lab staff positions. A
} } } } microbiologist with no em experience replaced me with 20 years of em
} } } } experience only because she lost her job due to downsizing and bumped
} } me out
} } } } of my position. The management had recently changed (lowered) all of the
} } } } standards/qualifications so this event could take place. The union was
} } } } never informed of this move, but that is another story.
} } } }
} } } } A supervisors position should not be part of the union. A smart manager
} } } } should realize that you cannot hire good help in this field into a
} } } } supervisor position without at least 7-10 years of EM experience.
} } } }
} } } } The minimum education should be at least a BS in a biological field
} } with at
} } } } least 5 years experience.
} } } }
} } } } Cheers,
} } } }
} } } } Ed Calomeni
} } } }
} } } }
} } } }
} } } } ----- Original Message -----
} } } } } From: "Dorota Wadowska" {wadowska-at-upei.ca}
} } } } To: {microscopy-at-sparc5.microscopy.com}
} } } } Sent: Tuesday, October 08, 2002 10:09 AM
} } } } Subject: EM job classificcation
} } } }
} } } }
} } } } }
} } ------------------------------------------------------------------------
} } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } America
} } } } } To Subscribe/Unsubscribe -- Send Email to
} } ListServer-at-MSA.Microscopy.Com
} } } } } On-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } }
} } -----------------------------------------------------------------------.
} } } } }
} } } } }
} } } } } Hi Listservers!
} } } } } I have a few questions that I would like to ask you If you feel you
} } } } } know the answer please reply. My university together with local
} } } } } union reclassified staff positions according to Aiken plan. Needless
} } } } } to say my position was downgraded.
} } } } } My questions are:
} } } } } 1. Can EM supervisory position be compared to any on campus
} } } } } staff positions (university has Department of Sciences and
} } } } } Veterinary Medecine)? I supervise small multi--user lab, we have 2
} } } } } TEM, and we do research, diagnostics, teaching (under- and
} } } } } graduate level) in the field of biological sciences. No fancy
} } staff like
} } } } } X-ray analysis or cryo.
} } } } } 2. What in your opinion is a necessary minimum experience
} } } } } required to perform this job ? I am talking about accepting the
} } } } } position and immediately getting into the stream of work.
} } } } } 3. What in your opinion is the minimum level of education that is
} } } } } required to perform this job?
} } } } }
} } } } } You can reply directly to me. Thanks in advance
} } } } } Frustrated
} } } } } Dorota
} } } } }
} } }



From daemon Mon Oct 14 16:05:36 2002



From: Mike Bode :      mb-at-Soft-Imaging.com
Date: Mon, 14 Oct 2002 14:35:58 -0600
Subject: RE: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I think, Peter and you are right with the statement that different light
paths are necessary to achive a true stereo effect. I am not so sure, that 2
separate objectives are really required. I know of at least one company (I
have no affiliation with or interest in that company) that makes a 3D
microscope and I think they are using only one objective lens
(http://www.edge-3d.com).

mike


Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: John W. Raffensperger, Jr. [mailto:johnr-at-helwigcp.com]
Sent: Monday, October 14, 2002 8:59 AM
To: 'by way of Ask-A-Microscopist'; Microscopy-at-sparc5.microscopy.com


Just because a microscope has two eyepieces, does not mean that it is a
"stereo" microscope.

A stereo microscope has not only two eyepieces, but also two objectives,
and two separate light paths. Because of having two slightly offset
images, you see a three dimensional image. Usually these are lower
powered systems, but still compound microscopes (multiple lens systems).

On a high power (relatively) microscope with a binocular head, you only
have one objective, and one light path from the specimen to the head.
The prism(s) in the head divide the light path to provide multiple
images, but they are all (three in the case of an extra camera port)
identical. This does not provide a 3-D image, it just makes it easier
on the eyes and brain to process the image because one eye is not
distracted.

I am not a scientist, and would be interested in any corrections to the
above.

John W. Raffensperger, Jr.
IS Manager
Helwig Carbon Products, Inc.


Email: gbarclay-at-trinidad.net
Name: Greg Barclay

Organization: University of the West Indies

Education: Graduate College

Location: St. Augustine, Trinidad, West Indies

Question: Light microscope terminology: We all know what a
stereomicroscope is and what a compound microscope is, but are they
not technically both the same thing?



From daemon Mon Oct 14 16:34:17 2002



From: Anna Johansson :      anna.johansson-at-bioseeker.com
Date: Mon, 14 Oct 2002 16:28:27 -0500 (CDT)
Subject: BSG BioNewsletter *Antifungals*Swedish Biopharma* 05#2002

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


BSG BioNewsLetter 05#2002 - BioSeeker Group's analysis and products eNewsletter

URL: http://www.bioseeker.com http://www.bsg-online.biz

-- CONTENTS --

1) Swedish Biopharma at Large - What Next?
2) New Report: 'The Next Generation Antifungals: Triazoles vs. Peptides'
3) Recently published BioSeeker Reports
4) Swedish Biotech Packages
5) Seek And You Will Find
6) Coming reports: HCV Therapy / Winning the Osteoporosis Therapy Race
7) Subscription
8) BioSeeker Online Store

====================================================================

1) SWEDISH BIOPHARMA INDUSTRY AT LARGE - WHAT NEXT?

The Swedish biopharma industry has more than doubled in the last six years, growing from 25 companies in 1996 to the current 52. Next to the medical device industry the biopharma industry is the largest biotech industry segment in Sweden. The predominant disease groups for R&D in Swedish biopharma are: Inflammation / autoimmune, metabolic, cancer, infectious, and neurological diseases. However, the pipeline of the industry is in its infancy.

Read the full version of this free, analytical BioNewsletter article at http://www.bioseeker.com/docs/topics/topic_swepharma.asp

====================================================================

2) THE NEXT GENERATION ANTIFUNGALS : TRIAZOLES VS PEPTIDES
- A New Report from BioSeeker Group

According to the Antifungal Work Group at NIH there is a "critical need for the development of antifungal agents". There will be tight competition for market shares when PfizerÕs patent on fluconazole (the active ingredient in Diflucan(TM)) expires in 2004. Both Bristol-Myers Squibb and Schering-Plough will try to compete for market shares. In this Highlight Report, we present new data on activites in the field of triazole and peptide antifungals. Topics considered in this report are :

- Research (spectrums, activities, side-effects, resistance and drug interactions)
- Patents
- Clinical trials (indications, status and activities)
- Partnerships
- Expected market launches and sales

For more information about this report, please visit http://www.bioseeker.com/index.asp?report=antifungal

====================================================================

3) RECENTLY PUBLISHED BIOSEEKER REPORTS

- Swedish Biopharma Industry - The Next Wave (released Sept, 2002)
http://www.swedishpharma.info

- Ion Channel Assays in the Drug Discovery Process (released June, 2002)
http://www.ion-channels-in-drug-discovery.info

- The Competitive Landscape of Cancer Gene Therapy (released May, 2002)
http://www.competitive-landscape-gene-therapy.info

- Commercializing on Tumor Antigens: Analytical Perspectives (released Feb, 2002)
http://www.commercializing-tumor-antigens.info

====================================================================

4) SWEDISH BIOTECH PACKAGES

a) Order the new report 'Swedish Biopharma Industry' together with the report 'Swedish Biotechnology Industry' (published Jan. 2001) for $1050 (save $200).

b) Obtain five free company profiles from the Swedish Biotech Corporate Index with any purchase of 'Swedish Biopharma Industry' (price $850 US).

c) Order 'Swedish Biotechnology Industry' together with the address list 'Complete Contact Data to Swedish Biotechnology Companies', updated continously and containing contact details to more than 320 Swedish Biotech companies, for $699 (save $350).

When ordering one of these packages, please place your order by fax and use reference codes SWEPACK-A, B or C. (Download orderform at www.bioseeker.com/docs/orderForm.pdf)

====================================================================

5) SEEK AND YOU WILL FIND

Identify and locate your potential customers, competitors, and partners by using BioSeeker Group's world-wide company search service. This service is based on BioSeeker's in-house company database built up during four years of operation and containing more than 10,000 life science companies globally.

We offer you this service in 3 steps:

Step 1. Search & Identify (399 USD)

Step 2. Locate & Validate (699 USD)

Step 3. Select & Build (individual price). Choose the companies of interest and BSG builds extensive profiles of those.

For more information about this service, please contact us at bioinfo-at-bioseeker.com

====================================================================

6) COMING REPORTS

- Molecular Approaches Towards New HCV Therapy

Treatment of HCV (Hepatitis C Virus) is a market with unmet medical needs, with relatively young people infected and a high risk for chronic disease. The coming decade will see exciting developments in HCV therapy thanks to breakthroughs in overcoming a major barrier to development: the lack of dedicated models suitable for validating candidate compounds in a low-cost, highthroughput format. BioSeeker Group has identified at least 115 companies with interest in various stages in the field of HCV therapy. Out of these 100, approximately 10 of the 'big pharma' are represented. Our coming report "Molecular Approaches Towards New HCV Therapy" analyses the field of new HCV therapeutics and assess possible outcomes for drugs and companies in this sector.

Please contact us at bioinfo-at-bioseeker.com for more information about this report.For further reading about HCV, please visit our free analytical article on this topic (http://www.bioseeker.com/docs/topics/topic_hcv.asp)

- Winning the Osteoporosis Therapy Race

Osteoporosis, deterioration of bone tissue and an increased risk of fractures, is considered by the World Health Organization to be second only to cardiovascular disease as a leading healthcare problem. According to other studies, sales of osteoporosis drugs in these markets totaled $3.4 billion in 2000, and are expected to reach $6.5 billion by 2005. A new class of osteoporosis therapies, designed to build bone mass in contrast to currently available anti-osteoporosis medicines are presently under study. These anabolic therapies consist of parathyroid hormone (FortÎoª and ALX-11) or fluoride to stimulate bone formation, but their safety and efficacy remain to be established. The first of these drugs to reach the market will probably be FortÎoª, known generically as teriparatide (recombinant parathyroid hormone).

Please contact us for more information about this report (bioinfo-at-bioseeker.com). For further reading about osteoporosis, please visit our free analytical article on this topic (http://www.bioseeker.com/docs/topics/topic_osteoporosis.asp)

====================================================================

7) SUBSCRIPTION

To continue to stay tuned with coming editions of BioSeeker Groups BioNewsLetter, please send an email to subscribe-at-bioseeker.com

============================================================================

8) BIOSEEKER ONLINE STORE

BioSeeker has now made it easy to order any of our products online, directly from our website. Visit www.bsg-online.biz to place an online order.

============================================================================

ADVERTISING
Do you have a message that you would like the biotech community to know? BioSeeker Group's Newsletter targets managers, scientists, venture capitalists, consultants and many more world wide. Contact us at sales-at-bioseeker.com for further information on how to advertise in BSG BioNewsletter.

COURTESY

If you would like to comment on our newsletter please send us an email to feedback-at-bioseeker.com

If you would like to be removed from our newsletter list, please reply to unsubscribe-at-bioseeker.com
















From daemon Mon Oct 14 17:14:49 2002



From: Walck, Scott D. :      walck-at-ppg.com
Date: Mon, 14 Oct 2002 18:01:56 -0400
Subject: Glue for TEM samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Isabel,
There are two epoxy glues that work well. Epo-Tek 353ND from Epoxy Technology is one. Gatan's G1 is equivalent to this. The other is M-bond 610. I like the Epo-Tek 353ND; mostly because it has a longer shelf life, but also because I believe that it is stronger. I believe that you can get thinner glue joints with the M-bond 610, though.

BTW, Your system is ideal for using the small angle cleavage technique. You can forgo glueing, dimpling, and ion milling altogether and make about 10 samples per hour. Check out the South Bay Technology web site and look at the Microcleave kit and examples that they have. A detailed description of the technique has been written up by John McCaffrey and myself in the number 4 proceedings book on TEM sample preparation from the MRS. John also has the original publication in the number 3 book from MRS.
-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8515 (fax)

"The opinions expressed are those of Scott D. Walck and not of PPG Industries, Inc. nor of any PPG-associated companies."



-----Original Message-----
} From: Isabel Nogueira [mailto:isabeln-at-popsrv.ist.utl.pt]
Sent: Monday, October 14, 2002 12:16 PM
To: Microscopy-at-sparc5.microscopy.com


Hello,

I need to prepare TEM cross-section samples of alternate SiO2 and TiO2
layers on a Si substrate. For that purpose I would appreciate suggestions on
glues, suitable for TEM work, that have a sputtering rate not too different
from my non homogeneous layers. The final sputtering will be on an Ion Mill
(4keV) using Ar+ ions.
All suggestions are welcome (including vendors) either to the list or to me
directly.
Thank you,

Isabel



Isabel Nogueira
Instituto Superior Técnico
Dep. Materiais
Avenida Rovisco Pais
1049-001 Lisboa
Portugal
tel.: +351 218418123
fax: +351 218418120
email: isabeln-at-popsrv.ist.utl.pt



From daemon Mon Oct 14 17:43:21 2002



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Mon, 14 Oct 2002 18:32:53 -0500
Subject: TEM cross-section glues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Isabel Nogueira wrote:
=========================================================
I need to prepare TEM cross-section samples of alternate SiO2 and TiO2
layers on a Si substrate. For that purpose I would appreciate suggestions on
glues, suitable for TEM work, that have a sputtering rate not too different
from my non homogeneous layers. The final sputtering will be on an Ion Mill
(4keV) using Ar+ ions.
All suggestions are welcome (including vendors) either to the list or to me
directly.
===========================================================
Several different glues have been mentioned on the listserver, one being M-
Bond™ 610, which is described on URL
http://www.2spi.com/catalog/spec_prep/glue.shtml

Disclaimer: SPI Supplies is a worldwide distributor for the M-Bond 610
product.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================




From daemon Mon Oct 14 19:56:04 2002



From: David Henriks :      henriks-at-southbaytech.com
Date: Mon, 14 Oct 2002 17:34:29 -0700
Subject: Re: Glue for TEM samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Isabel:

Some very popular "glues" used for bonding TEM samples for cross section
analysis are:

Product Name: Epo-Tek 353ND

Epoxy Technology
TEL: 978-667-3805
FAX: 978-663-9782
www.epotek.com

Product Name: Loctite 460
http://www.loctite-europe.com/es/ (Website in Spain)

Loctite Corp
1001 Trout Brook Crossing
Rocky Hill, CT 06067 USA
TEL: 860.571.510
www.loctite.com

Product Name: M-Bond 610

Vishay Measurements Group, Inc.
P.O. Box 27777
Raleigh, North Carolina 27611 USA

TEL: 919-365-3800
FAX: 919-365-3945
measurementsgroup-at-vishay.com

I hope this helps.

Best regards-

David

--
David Henriks
Vice President

South Bay Technology, Inc.
1120 Via Callejon
San Clemente, CA 92673 USA

TEL: +1-949-492-2600
Toll-free in the USA: +1-800-728-2233
FAX: +1-949-492-1499

email: henriks-at-southbaytech.com

Celebrating 38 years of providing Materials Processing Solutions for
Metallogaphy, Crystallography and Electron Microscopy.

Please visit us online at www.southbaytech.com.
Isabel Nogueira wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hello,
}
} I need to prepare TEM cross-section samples of alternate SiO2 and TiO2
} layers on a Si substrate. For that purpose I would appreciate suggestions on
} glues, suitable for TEM work, that have a sputtering rate not too different
} from my non homogeneous layers. The final sputtering will be on an Ion Mill
} (4keV) using Ar+ ions.
} All suggestions are welcome (including vendors) either to the list or to me
} directly.
} Thank you,
}
} Isabel
}
} Isabel Nogueira
} Instituto Superior Técnico
} Dep. Materiais
} Avenida Rovisco Pais
} 1049-001 Lisboa
} Portugal
} tel.: +351 218418123
} fax: +351 218418120
} email: isabeln-at-popsrv.ist.utl.pt



The information contained in this message and any attachments is
privileged and
confidential. This message is intended for the individual or entity
addressed.
If you are not the intended recipient, please do not read, copy or
disclose this
communication. Notify the sender of the mistake by calling
+1-949-492-2600 and
delete this message from your system.


From daemon Mon Oct 14 22:26:34 2002



From: Gordon Couger :      gcouger-at-couger.com
Date: Mon, 14 Oct 2002 22:17:02 -0500
Subject: Re: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Almost all modern stereo microscopes use a single objective lens that have
two light paths of varying complexity viewing the specimen at different view
points through a single objective.

The light paths are not parallel but view the subject from a slightly
different angle and present them to the eye at an angle compatible with the
brains processing methods to make it appear as a 3D image. How the light
paths are handled between the subject and the eye may have them at any
relation to each other but when they reach the eye they are at an angle that
make them appear to be 3D images.

Viewing a subject at an angle though a common objective introduces some
distortion but less than viewing the same subject through two completely
separate microscopes that are an angle to the subject.

If you have doubts about this find a AO Cycloptic with misaligned lenses in
the cylinder that change the magnification and see some of the weird ways an
image can behave when it is presented to the eyes incorrectly.

Gordon

Gordon Couger
Stillwater, OK
www.couger.com/gcouger


} From: "Peter Tomic" {PTomic-at-anadigics.com}
:
: I must jump in on this one. A real "stereo" microscope has two parallel
: optical paths. That is, two objective lenses. No microscope with one
: objective lens can be "stereo." Albeit there may be two eyepieces but
: without two separate optical paths the images in both eyepieces must
: necessarily be identical. It's like listening to stereo sound but if the
: left and right channels are identical, hence no depth or direction to the
: sound.
:
: Inverted or non-inverted images do not define a compound microscope since
: many are made with an "image erector" which is simply an additional lens
to
: reverse the image so you view the sample as it is with the unaided eye.
: Most "micromanipulator" systems have image erectors on them for that
: purpose. It helps one to not have to re-wire one's neurons to interpret
: left as right when having to use hand manipulation under the microscope.
I
: have found that wiring and re-wiring neurons problematic.
:
: Peter
: Anadigics, Inc.
:
: -----Original Message-----
: } From: qualityimages [mailto:qualityimages-at-netrax.net]
: Sent: Saturday, October 12, 2002 6:25 PM
: To: by way of Ask-A-Microscopist; Microscopy
: Subject: Re: Ask-A-Microscopist:LM terminology
:
:
: ------------------------------------------------------------------------
: The Microscopy ListServer -- Sponsor: The Microscopy Society of America
: To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
: On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
: -----------------------------------------------------------------------.
:
:
: It seems to me that there jas always been one other difference between
: the two. Most truly binocular (stereo) microscopes do not have an
: inverted image, whereas most "compound" microscopes have an inverted
image.
:
: Ken Converse
: owner
: Quality Images
: third party SEM service
: Delta, PA
:
: by way of Ask-A-Microscopist wrote:
:
: } ------------------------------------------------------------------------
: } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
: } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
: } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
: } -----------------------------------------------------------------------.
: }
: }
: } Below is the result of your feedback form (NJZFM-ultra-55). It was
: } submitted by (gbarclay-at-trinidad.net) from
: } http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday,
: } October 11, 2002 at 23:40:22
: }
: --------------------------------------------------------------------------
-
: }
: }
: } Email: gbarclay-at-trinidad.net
: } Name: Greg Barclay
: }
: } Organization: University of the West Indies
: }
: } Education: Graduate College
: }
: } Location: St. Augustine, Trinidad, West Indies
: }
: } Question: Light microscope terminology: We all know what a
: } stereomicroscope is and what a compound microscope is, but are they
: } not technically both the same thing? Should not the distinction be
: } between the "simple microscope," basically a highly corrected doublet
: } or triplet lens, forming the Dowland type of microscope, and any kind
: } of microscope with a separate eyepiece and objective lens? If so, then
: } both the stereo and binocular (or monocular) microscopes are compond
: } microscopes. I have been referring to "stereomicroscopes and binocular
: } microscopes" in our Life Sciences Dept, and having some arguments with
: } those who refer "stereomicroscopes and compound microscopes," over
: } what is really what.
: }
: }
: --------------------------------------------------------------------------
-
: }
: }
: }
:
:
:
:
:
:



From daemon Tue Oct 15 07:15:36 2002



From: Jim Quinn :      jquinn-at-www.matscieng.sunysb.edu
Date: Tue, 15 Oct 2002 08:04:24 -0400
Subject: RE: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Mike -

re: stereo/terminology

I agree, my Lynx is stereo with only one objective.

http://www.visioneng.com/

Jim

} From Microscopy-request-at-sparc5.microscopy.com Tue Oct 15 03:54:02 2002
} From: Mike Bode {mb-at-Soft-Imaging.com}
} To: "'Microscopy-at-MSA.Microscopy.Com'" {Microscopy-at-sparc5.microscopy.com}
} Subject: RE: Ask-A-Microscopist:LM terminology
} Date: Mon, 14 Oct 2002 14:35:58 -0600
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I think, Peter and you are right with the statement that different light
} paths are necessary to achive a true stereo effect. I am not so sure, that 2
} separate objectives are really required. I know of at least one company (I
} have no affiliation with or interest in that company) that makes a 3D
} microscope and I think they are using only one objective lens
} (http://www.edge-3d.com).
}
} mike
}
}
} Michael Bode, Ph.D.
} Soft Imaging System Corp.
} 12596 West Bayaud Avenue
} Suite 300
} Lakewood, CO 80228
} ===================================
} phone: (888) FIND SIS
} (303) 234-9270
} fax: (303) 234-9271
} email: mailto:info-at-soft-imaging.com
} web: http://www.soft-imaging.com
} ===================================
}
}
}
} -----Original Message-----
} } From: John W. Raffensperger, Jr. [mailto:johnr-at-helwigcp.com]
} Sent: Monday, October 14, 2002 8:59 AM
} To: 'by way of Ask-A-Microscopist'; Microscopy-at-sparc5.microscopy.com
} Subject: RE: Ask-A-Microscopist:LM terminology
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Just because a microscope has two eyepieces, does not mean that it is a
} "stereo" microscope.
}
} A stereo microscope has not only two eyepieces, but also two objectives,
} and two separate light paths. Because of having two slightly offset
} images, you see a three dimensional image. Usually these are lower
} powered systems, but still compound microscopes (multiple lens systems).
}
} On a high power (relatively) microscope with a binocular head, you only
} have one objective, and one light path from the specimen to the head.
} The prism(s) in the head divide the light path to provide multiple
} images, but they are all (three in the case of an extra camera port)
} identical. This does not provide a 3-D image, it just makes it easier
} on the eyes and brain to process the image because one eye is not
} distracted.
}
} I am not a scientist, and would be interested in any corrections to the
} above.
}
} John W. Raffensperger, Jr.
} IS Manager
} Helwig Carbon Products, Inc.
}
}
} Email: gbarclay-at-trinidad.net
} Name: Greg Barclay
}
} Organization: University of the West Indies
}
} Education: Graduate College
}
} Location: St. Augustine, Trinidad, West Indies
}
} Question: Light microscope terminology: We all know what a
} stereomicroscope is and what a compound microscope is, but are they
} not technically both the same thing?
}
}


From daemon Tue Oct 15 07:32:27 2002



From: Scott Whittaker :      Whittaker.scott-at-nmnh.si.edu
Date: Tue, 15 Oct 2002 08:25:39 -0400
Subject: Re: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


To enhance contrast a bit we kept stock bottles of 50% and 70% etoh with
1-2% UA adding contrast while dehydrating. As long as tissue pieces were
small this worked well. Solution was drawn from the top or filtered to
prevent the precip that collected in the bottom. Time was doubled in these
two steps. Obviously if extraction or glycogen was important then this was
not done.

Scott Whittaker
Laboratories of Analytical Biology
Smithsonian Institution
National Museum of Natural History
PO Box 37012 MRC104
Washington DC 20013-7012
202-357-1651


} } } mohammed y abdulrawoof / f40z006 75760 {mdyousuf-at-KSU.EDU.SA} 10/10/02
12:35PM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Dear listers,
Please advise as to the usefulness/advantage of doing enblock staining of
biological tissues as against doing heavy-metal staining of grids with
thin sections. The tissues that I am about to process are for diagnosis
(pathology).

Thanks
M. Yousuf Abdul-Rawoof




From daemon Tue Oct 15 07:44:50 2002



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Tue, 15 Oct 2002 13:38:02 +0100 (GMT Daylight Time)
Subject: Re: ESEM vs. gold coated SEM images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


ESEM images of bacteria are different from gold coated
images of either air dried or CPD bacteria.

Firstly many bacteria produce slime. In air dried
samples this collapses. In CPD samples the solvents
probably remove it or condense it. In ESEM the slime is
retained so that one often does not see naked bacteria at
all. This conflict with expectation can lead to
disappointment!

Secondly, gold gives a good SE yield from the surface. In
ESEM (tungsten filament) bacteria yield less SE and from a
greater depth. This gives a "softer" image of bacteria than
one is used to.

Don't worry about coating units being outmoded. We showed
school visitors 7 samples today. All were gold coated!
The samples are easier to image.

With many dry materials samples it is easier to gold coat
and then image in ESEM. I know this is cheating ....

We are still using CPD for projects comparing preparation
methods.

Dave


On Thu, 10 Oct 2002 12:40:14 -0500 "Garber, Charles A."
{cgarber-at-2spi.com} wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} -- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --
}
} David Patton wrote:
} ==========================================================
} Disclaimer: ESEM images of bacteria have a different quality to those that
} } have been dried and gold coated; so warn your user to avoid
} disappointment.
} ===========================================================
}
} Are you comparing ESEM images with a) air dried and gold coated or b)
} critical point dried and gold coated?
}
} It would be obvious that comparing air dried would indeed show a difference
} but if you really were comparing it to critical point dried samples, could
} you elaborate a bit on the nature of those differences? And would you
} have any comment about differences in contrast comparing ESEM vs. gold
} coated images.
}
}
} Disclaimer: SPI Supplies offers critical point dryers and sputter coaters so
} I guess we would be a bit nervous of any technology that should evolve that
} would render obsolete at the same time both CPD units and sputter coaters!
}
} Chuck
} ============================================
}
} Charles A. Garber, Ph. D. Ph: 1-610-436-5400
} President 1-800-2424-SPI
} SPI SUPPLIES FAX: 1-610-436-5755
} PO BOX 656 e-mail:cgarber-at-2spi.com
} West Chester, PA 19381-0656 USA
} Cust.Service: spi2spi-at-2spi.com
}
} Look for us!
} ########################
} WWW: http://www.2spi.com
} ########################
} ============================================
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



From daemon Tue Oct 15 08:56:18 2002



From: George_Munzing-at-engelhard.com
Date: Tue, 15 Oct 2002 09:22:47 -0400
Subject: Need a high temperature epoxy for microtoming

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello fellow microscopists,

I need to do some high temperature TEM studies and am looking for a
suitable mounting resin. I will be embedding and microtoming powder
materials and would like to go as high in temperature as possible. The
materials themselves can withstand the temperatures but I doubt my
conventional epoxy would survive.

Is there any silicone epoxy or other suitable material designed to
withstand high temperatures (maybe as high as 600-800C) for a short time
for embedding?

Any suggestions you can offer are greatly appreciated.

Thanks in advance!

George R. Munzing Jr.
Engelhard Corporation
Strategic Technologies Group
25 Middlesex/Essex Turnpike
Iselin, NJ 08830
Tel: 732-205-7030
FAX: 732-205-5300



From daemon Tue Oct 15 10:21:49 2002



From: Shane Roberts :      roberts-at-southbaytech.com
Date: Tue, 15 Oct 2002 08:02:58 -0700
Subject: Glue for TEM samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


There is another 'glue' that allows you to tailor the ion milling rate. The
product is called Araldit or LEIT-C. It is a carbon based glue that starts
as a powder, transitions into a viscous liquid, then polymerizes shortly
thereafter. The advantage is the glue mills at a slightly lower rate than
normal epoxies, and you can increase the carbon content to further reduce
the milling rate. South Bay Technology does supply this type of product.
Please let me know if you would like further information.

Best Regards,
----------------------------------
Shane Roberts
Applications Engineer
South Bay Technology,Inc.
'Materials Processing Solutions'
www.southbaytech.com
phone: 949.492.2600
fax: 949.492.1499
email: roberts-at-southbaytech.com
-----------------------------------

-----Original Message-----
} From: Isabel Nogueira [mailto:isabeln-at-popsrv.ist.utl.pt]
Sent: Monday, October 14, 2002 9:16 AM
To: Microscopy-at-sparc5.microscopy.com


Hello,

I need to prepare TEM cross-section samples of alternate SiO2 and TiO2
layers on a Si substrate. For that purpose I would appreciate suggestions on
glues, suitable for TEM work, that have a sputtering rate not too different
from my non homogeneous layers. The final sputtering will be on an Ion Mill
(4keV) using Ar+ ions.
All suggestions are welcome (including vendors) either to the list or to me
directly.
Thank you,

Isabel



Isabel Nogueira
Instituto Superior Técnico
Dep. Materiais
Avenida Rovisco Pais
1049-001 Lisboa
Portugal
tel.: +351 218418123
fax: +351 218418120
email: isabeln-at-popsrv.ist.utl.pt





From daemon Tue Oct 15 10:31:04 2002



From: mohammed y abdulrawoof / f40z006 75760 :      mdyousuf-at-KSU.EDU.SA
Date: Tue, 15 Oct 2002 18:17:07 -0300 (GMT)
Subject: Thanks: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear All who replied,
Using both en block staining for the tissues followed by double staining
of grids with UA and Lead citrate seems to be a good idea. Anyway, to get
a personal insight and experience, I will try fixation with and without
en-block staining.
Thanks again.
M. Yousuf.




From daemon Tue Oct 15 15:27:40 2002



From: Geoff Williams :      willi1gl-at-cmich.edu
Date: Tue, 15 Oct 2002 16:14:45 -0400
Subject: RE: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I saw the first post this weekend and am jut now getting to it - doesn't
look like the answer is really been quite addressed yet... and yes even
at the risk of starting something . . . . ;)

What a wonderful question. Another semantic, logical, debatable but
ultimately one answer question!

Compound microscopes are as you say anything with a compound series of
lenses. Yes a 'Stereo' microscope should be called a compound stereo
microscope (or some variation). The distinction is that (I feel) the
"compound" microscope should be call a transmitted (or epi-illuminated)
light microscope. But then that would demonstrate an inclination of an
electron microscope bias.

There is nothing wrong with the term stereomicroscope, but the issue
with changing the compound microscope's name is more difficult than
worth it. It is much like trying to get the rest of the US to start
using the metric system. The terminology is too entrenched, but it
works, so why fix it?

The term binocular then becomes similarly confusing as Stereo
microscopes are also 'binocular' the fixed definition isn't much better
than the 'compound' in that it has the same issues. But then as
addressed before binocular compound 'transmitted' microscopes are not
'stereo' microscopes.

Greg, call 'em what you want - as long as your definition is made clear
to your audience and the distinctions made in the discussion on the list
server is kept in mind.

Geoff Williams
Microscopy Facility Supervisor


} -----Original Message-----
} From: by way of Ask-A-Microscopist [mailto:gbarclay-at-trinidad.net]
} Sent: Saturday, October 12, 2002 2:48 PM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: Ask-A-Microscopist:LM terminology
}
}
------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
-----------------------------------------------------------------------.
}
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (gbarclay-at-trinidad.net) from
} http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday,
} October 11, 2002 at 23:40:22
}
------------------------------------------------------------------------
--
} -
}
} Email: gbarclay-at-trinidad.net
} Name: Greg Barclay
}
} Organization: University of the West Indies
}
} Education: Graduate College
}
} Location: St. Augustine, Trinidad, West Indies
}
} Question: Light microscope terminology: We all know what a
} stereomicroscope is and what a compound microscope is, but are they
} not technically both the same thing? Should not the distinction be
} between the "simple microscope," basically a highly corrected doublet
} or triplet lens, forming the Dowland type of microscope, and any kind
} of microscope with a separate eyepiece and objective lens? If so,
} then both the stereo and binocular (or monocular) microscopes are
} compond microscopes. I have been referring to "stereomicroscopes and
} binocular microscopes" in our Life Sciences Dept, and having some
} arguments with those who refer "stereomicroscopes and compound
} microscopes," over what is really what.
}
}
------------------------------------------------------------------------
--
} -



From daemon Wed Oct 16 02:09:08 2002



From: wang lan :      wins-at-public.xa.sn.cn
Date: Wed, 16 Oct 2002 14:47:15 +0800
Subject: optical products binocurlas, microscopes, telescopes, riflescopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


We make new optical instruments for importers:

www.winsbinoculars.com

Regards,


Wentao


From daemon Wed Oct 16 02:09:13 2002



From: wang lan :      wins-at-public.xa.sn.cn
Date: Wed, 16 Oct 2002 14:53:13 +0800
Subject: optical products binocurlas, microscopes, telescopes, riflescopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


We make new optical instruments for importers:

www.winsbinoculars.com

Regards,


Wentao


From daemon Wed Oct 16 08:23:01 2002



From: Richard Beanland :      richard.beanland-at-bookham.com
Date: Wed, 16 Oct 2002 14:11:40 +0100
Subject: WORM drives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello everyone,
although this is strictly not a microscopy question, I'm posting
this in the hope that someone has been through this process already and can
offer some advice.

Our LEO 360FE SEM computer saves images on a Panasonic LF-7300 WORM drive.
It doesn't seem feasible to connect the computer to our intranet directly
(too old/difficult!) but a relatively straighforward way to get at the
images is to install an identical WORM drive on a PC which is networked.
However I imagine that some kind of software is needed to make the drive
functional under Windows NT. My internet searches for drivers have drawn
blanks so far. Has anyone done this and can give me some advice?
We are asking LEO for some support on this but independent advice is
always very valuable.

Many thanks,

Richard


_______________________________
Richard Beanland
Analytical Services
Bookham Technology plc
Caswell,
Towcester,
Northamptonshire NN12 8EQ
UK
Tel: +44 (0) 1327 356362
Fax: +44 (0) 1327 356398
http://www.bookham.com



=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.

No part of this message can be considered a request for goods or
services.
=======================================================================
Any questions about Bookham's E-Mail service should be directed to postmaster-at-bookham.com.



From daemon Wed Oct 16 13:08:48 2002



From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Wed, 16 Oct 2002 07:54:23 -1000 (HST)
Subject: Foil covered SEM stubs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi, all-

Here is a colleague's method for using aluminum foil on SEM stubs for a
smooth background:

1. "Reynolds" seems smoothest and "regular" seems easiest to work with.

2. Burnishing. I've tried various things and they usually end up
scratching the foil. To make sure it is absolutely smooth flat, I
carefully cut a strip that is a bit wider and longer than a stub.
Place it on a clean surface, cover it with something very smooth
(weighing paper works well) and work over it with a finger tip to get
it absolutely as flat as possible. Picking it up by the very edge
with tweezers helps avoid wrinkles.

3. The smooth, flat piece is then transferred to a piece of
double-stick scotch tape and once more pressed to the tape to get a
very flat surface with no bubbles between foil and tape. This should
be done on glass or a surface from which tape plus foil can then be
carefully lifted.

4. The tape plus foil can then be applied to the stub - if it still
looks nice, proceed to step five. If wrinkles or bumps have appeared,
pull it up and try again. (You can put the tape directly on the stub
and then apply the foil - However, I have found that it works best
for me to have a nicely bonded duo of smooth tape and foil first)

5. Using weighing paper, again, press tape plus foil gently to the stub.

6. Finally, use a single-edge razor blade to trim off the excess foil
and tape around the edge of the stub. You can apply a couple of teeny
dabs of silver paste at several edge points if you want to make
absolutely certain the foil is well connected to the stub. I have
gotten into the habit of making the strip of foil plus tape slightly
narrower than stub diameter so that there is a small crescent of stub
above and below the strip, simply as a way of telling top and bottom
at a glance.

7. Additional note. For mounting an object with a convex (like my
little larval shells) or irregular lower surface, aluminum foil can
be gouged with a minuten insect pin before applying silver paste or
paint to help achieve better contact and bonding.

As with everything, it just takes a bit of experimentation. . . .


I hope this helps!

Aloha,
Tina



From daemon Wed Oct 16 13:15:00 2002



From: Karl Garsha :      garsha-at-itg.uiuc.edu
Date: Wed, 16 Oct 2002 13:08:25 -0500
Subject: Re: Ask-A-Microscopist:epi-fluorescence

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Greetings Sundeep,
If you try entering "polarized fluorescence microscopy" into the Google
search engine you will get some hits. Using a polarized light source in
conjunction with a crossed polarizing filter which can be rotated
(analyzing filter) is a technique used in different approaches to
fluorescence microscopy to observe molecular rotations in protein folding
or in TIRF applications. It can be used in conjunction with FRET type
experiments. I believe the book entitled "Fluorescence Microscopy"
published by Bios Scientific may have a short section which discusses
polarized fluorescence microscopy.
Regards,
Karl G.

At 06:11 PM 10/10/2002 -0500, sundeep.bhandari-at-nikon.co.uk wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_______________________________________________
Karl Garsha
Light Microscopy Specialist
Imaging Technology Group
Beckman Institute for Advanced Science and Technology
University of Illiniois at Urbana-Champaign
405 North Mathews Avenue
Urbana, IL 61801
Room B650J
Tel: (217) 244-6292
Fax: (217) 244-6219
www.itg.uiuc.edu



From daemon Wed Oct 16 13:43:17 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Wed, 16 Oct 2002 14:41:15 -0400
Subject: RE: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Mike & Co.

We also have an Olympus Model SZH10 that has only one "objective" but it's
not really an objective lens [1X to 80X total mag. for microscope]. It's
simply a mag. doubler in front of the pair of objective lenses. If one
turns the microscope upside down and removes this mag. doubler, they would
see two objectives and there are two optical paths all the way up to the
eyepieces. Ain't no way to have stereo with one optical path. I've never
seen a "metallurgical" "compound" microscope that has mags. up to 1 KX. with
two distinct optical paths.



Peter

-----Original Message-----
} From: Mike Bode [mailto:mb-at-Soft-Imaging.com]
Sent: Monday, October 14, 2002 4:36 PM
To: 'Microscopy-at-MSA.Microscopy.Com'


I think, Peter and you are right with the statement that different light
paths are necessary to achive a true stereo effect. I am not so sure, that 2
separate objectives are really required. I know of at least one company (I
have no affiliation with or interest in that company) that makes a 3D
microscope and I think they are using only one objective lens
(http://www.edge-3d.com).

mike


Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: John W. Raffensperger, Jr. [mailto:johnr-at-helwigcp.com]
Sent: Monday, October 14, 2002 8:59 AM
To: 'by way of Ask-A-Microscopist'; Microscopy-at-sparc5.microscopy.com


Just because a microscope has two eyepieces, does not mean that it is a
"stereo" microscope.

A stereo microscope has not only two eyepieces, but also two objectives,
and two separate light paths. Because of having two slightly offset
images, you see a three dimensional image. Usually these are lower
powered systems, but still compound microscopes (multiple lens systems).

On a high power (relatively) microscope with a binocular head, you only
have one objective, and one light path from the specimen to the head.
The prism(s) in the head divide the light path to provide multiple
images, but they are all (three in the case of an extra camera port)
identical. This does not provide a 3-D image, it just makes it easier
on the eyes and brain to process the image because one eye is not
distracted.

I am not a scientist, and would be interested in any corrections to the
above.

John W. Raffensperger, Jr.
IS Manager
Helwig Carbon Products, Inc.


Email: gbarclay-at-trinidad.net
Name: Greg Barclay

Organization: University of the West Indies

Education: Graduate College

Location: St. Augustine, Trinidad, West Indies

Question: Light microscope terminology: We all know what a
stereomicroscope is and what a compound microscope is, but are they
not technically both the same thing?



From daemon Wed Oct 16 13:52:43 2002



From: akc-at-umich.edu
Date: Wed, 16 Oct 2002 14:46:01 -0400
Subject: RE: Ask-A-Microscopist:LM terminology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


In the history of microscopy, the term "compound microscope" was used in
contrast to a "simple microscope." A simple microscope had one lens, while
a compound microscope had two or more lenses (for example, an eyepiece).
Thus in the 1600s, Leeuwenhoek used a simple microscope (one lens of rather
high magnification), while Hooke did his work with a compound microscope
(which had an eyepiece). Compound microscopes were much more convenient to
use and were easier on the eyes. On the other hand, simple microscopes
were less affected by chromatic abberation, which allowed Leeuwenhoek to
see smaller objects (bacterial, sperm, and so forth) than could be seen
with the compound microscopes of the time. Not until the 1820s, with the
advent of achromatic lenses for compound microscopes, could microscopists
improve much on the detail of Leeuwenhoek's observations.

Kent

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
A. Kent Christensen, Professor Emeritus
Department of Cell and Developmental Biology, Medical Science II Building
University of Michigan Medical School, Ann Arbor, MI 48109-0616
Tel (work) (734) 763-1287, Fax (work) (734) 763-1166
akc-at-umich.edu http://www.umich.edu/~akc/
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

--On Tuesday, October 15, 2002 4:14 PM -0400 Geoff Williams
{willi1gl-at-cmich.edu} wrote:

} I saw the first post this weekend and am jut now getting to it - doesn't
} look like the answer is really been quite addressed yet... and yes even
} at the risk of starting something . . . . ;)
}
} What a wonderful question. Another semantic, logical, debatable but
} ultimately one answer question!
}
} Compound microscopes are as you say anything with a compound series of
} lenses. Yes a 'Stereo' microscope should be called a compound stereo
} microscope (or some variation). The distinction is that (I feel) the
} "compound" microscope should be call a transmitted (or epi-illuminated)
} light microscope. But then that would demonstrate an inclination of an
} electron microscope bias.
}
} There is nothing wrong with the term stereomicroscope, but the issue
} with changing the compound microscope's name is more difficult than
} worth it. It is much like trying to get the rest of the US to start
} using the metric system. The terminology is too entrenched, but it
} works, so why fix it?
}
} The term binocular then becomes similarly confusing as Stereo
} microscopes are also 'binocular' the fixed definition isn't much better
} than the 'compound' in that it has the same issues. But then as
} addressed before binocular compound 'transmitted' microscopes are not
} 'stereo' microscopes.
}
} Greg, call 'em what you want - as long as your definition is made clear
} to your audience and the distinctions made in the discussion on the list
} server is kept in mind.
}
} Geoff Williams
} Microscopy Facility Supervisor
}
}
} } -----Original Message-----
} } From: by way of Ask-A-Microscopist [mailto:gbarclay-at-trinidad.net]
} } Sent: Saturday, October 12, 2002 2:48 PM
} } To: Microscopy-at-sparc5.microscopy.com
} } Subject: Ask-A-Microscopist:LM terminology
} }
} }
} ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} -----------------------------------------------------------------------.
} }
} }
} } Below is the result of your feedback form (NJZFM-ultra-55). It was
} } submitted by (gbarclay-at-trinidad.net) from
} } http://www.msa.microscopy.com/Ask-A-Microscopist.html on Friday,
} } October 11, 2002 at 23:40:22
} }
} ------------------------------------------------------------------------
} --
} } -
} }
} } Email: gbarclay-at-trinidad.net
} } Name: Greg Barclay
} }
} } Organization: University of the West Indies
} }
} } Education: Graduate College
} }
} } Location: St. Augustine, Trinidad, West Indies
} }
} } Question: Light microscope terminology: We all know what a
} } stereomicroscope is and what a compound microscope is, but are they
} } not technically both the same thing? Should not the distinction be
} } between the "simple microscope," basically a highly corrected doublet
} } or triplet lens, forming the Dowland type of microscope, and any kind
} } of microscope with a separate eyepiece and objective lens? If so,
} } then both the stereo and binocular (or monocular) microscopes are
} } compond microscopes. I have been referring to "stereomicroscopes and
} } binocular microscopes" in our Life Sciences Dept, and having some
} } arguments with those who refer "stereomicroscopes and compound
} } microscopes," over what is really what.
} }
} }
} ------------------------------------------------------------------------
} --
} } -
}








From daemon Wed Oct 16 15:14:00 2002



From: Jensen, Karen :      Karen_Jensen-at-urmc.rochester.edu
Date: Wed, 16 Oct 2002 15:56:44 -0400
Subject: Brain Fixative

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers:

Does anyone have a recipe for mouse brain perfusion and fixation? I am
trying to get the best brain preservation in mice. I trying to prevent
neurons from swelling and I need to look at the Golgi apparatus. How about
calcium chloride as an additive?

Thanks!

Karen

Karen L. Bentley, M.S.(previously Jensen)
Associate Scientist & Project Manager
Electron Microscope Research Core
University of Rochester Medical Center
Rochester, NY 14642
585-275-1954



From daemon Wed Oct 16 18:23:03 2002



From: White, Woody N. :      nwwhite-at-mcdermott.com
Date: Wed, 16 Oct 2002 18:07:31 -0500
Subject: WORM drives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


If your drive is the old WORM technology, good luck finding one! I haven't
seen one for sale in years.

What computer does the 360 use? I have forgotten... Even if it is a DEC
(perhaps especially) it would be easier to send the data to a newer PC for
storage on hard drives and CD-Rs.

Regards, Woody

Woody White
McDermott Technology Inc.
McD: http://www.mtiresearch.com/
Mine: http://woody.white.home.att.net


-----Original Message-----
} From: Richard Beanland [mailto:richard.beanland-at-bookham.com]
Sent: Wednesday, October 16, 2002 9:12 AM
To: 'Microscopy-at-MSA.Microscopy.Com'


Hello everyone,
although this is strictly not a microscopy question, I'm posting
this in the hope that someone has been through this process already and can
offer some advice.

Our LEO 360FE SEM computer saves images on a Panasonic LF-7300 WORM drive.
It doesn't seem feasible to connect the computer to our intranet directly
(too old/difficult!) but a relatively straighforward way to get at the
images is to install an identical WORM drive on a PC which is networked.
However I imagine that some kind of software is needed to make the drive
functional under Windows NT. My internet searches for drivers have drawn
blanks so far. Has anyone done this and can give me some advice?
We are asking LEO for some support on this but independent advice is
always very valuable.

Many thanks,

Richard


_______________________________
Richard Beanland
Analytical Services
Bookham Technology plc
Caswell,
Towcester,
Northamptonshire NN12 8EQ
UK
Tel: +44 (0) 1327 356362
Fax: +44 (0) 1327 356398
http://www.bookham.com



=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.

No part of this message can be considered a request for goods or
services.
=======================================================================
Any questions about Bookham's E-Mail service should be directed to
postmaster-at-bookham.com.



From daemon Thu Oct 17 03:17:23 2002



From: Peter Heimann :      peter.heimann-at-uni-bielefeld.de
Date: Thu, 17 Oct 2002 09:48:24 +0200
Subject: Re: Brain Fixative

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Karen:

from my experience with mouse (CNS; spinal chord, muscle, testis
and other tissue) I would strongly recommend a mixture of 2%
paraformaldehyde / 4% glutaraldehyde (made from "the good one" prepared
after"Anderson")and a short (!) pre-perfuse (as you name it) to flush
out all blood cells, which otherwise regularly clog and occlude the fine
capillaries.
However, though simple saline flush can be sufficient, I routinely
use a mixture of Heparin (anti-coagulant), Procain (keeps the blood
vessels "open"), PVP (polyvinylpyrrolidon, MW-class 30 K; for
osmotic pressure) in aqua dest. and pH-adjustment.
Method published originally in a wonderful perfusion-technique
paper by Forssmann, WG et al (1977) in Anat. Rec. 188: p 307-
314. Just follow their technical procedure exactly (except that in the
small mouse you have to use a 26 gauge butterfly teflon-needle) and
you`ll get excellent fixation.
good luck!
peter
**********************************
peter.heimann-at-uni-bielefeld.de
Dr. Peter Heimann
Developmental Biology & Molecular Pathology; W7-107
University of Bielefeld
D 33501 Bielefeld / Germany
phone: xx49(0)521-106-5628 / 5627
FAX : " " - 5654
www.uni-bielefeld.de/biologie/Entwicklungsbiologie/
www.uni-bielefeld.de/SFB549
***********************************



From daemon Thu Oct 17 03:22:06 2002



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Thu, 17 Oct 2002 01:21:08 -0700
Subject: Re: Brain Fixative

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Karen,

I perfuse mouse with 1x PBS (20 mM Na-phosphate, 150 mM NaCl, PH 7.4) 15
ml. 2 ml/min, room temp, then same amount/speed 0.1% GA + 2% formaldhyde in
1x PBS, room temp. Brain as quick as possible removed from the scull on ice
and sliced 0.5 mm thick. Slices individually fixed in 1% GA/2%
formaldehyde/1x PBS at +4oC overnight.

I did not check Golgi but mitochondria and other stuff is OK. Sergey

P.S. I did not try Ca+2, but it's good combination with formaldehyde for
brain.

P.P.S. All chemicals are EM grade and fixers prepared just before
use. Slicing is critical. Speed of preparation is VERY critical. The way
you kill animal is also critical. Godd luck!

At 03:56 PM 10/16/02 -0400, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From daemon Thu Oct 17 05:00:39 2002



From: Albert Coritz :      cactusgrower-at-earthlink.net
Date: Thu, 17 Oct 2002 05:52:15 -0400
Subject: Help with Decalcification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Everyone:

Just passing this along for Matt who is not on the listserver. Please
email him directly with any ideas.

Many Thanks!

Al Coritz
Electron Microscopy Sciences

I was wondering if you knew any good protocols for decalcifying the
mouse skull. We haven't done too much of it, but when we have, we've
used 3%HCl and 15%NaCl.
As always, we're looking for a way to keep shrinkage of the brain to a
minimum. Do you have any ideas? Any help would be appreciated.
Thanks a lot.
Matt

Matt McElwee
Research Specialist
Department of Surgery
K4/617 Clinical Science Center
600 Highland Avenue
Madison, WI 53792
mcelwee-at-surgery.wisc.edu



From daemon Thu Oct 17 07:02:39 2002



From: Everett Ramer :      eramer-at-cellomics.com
Date: Thu, 17 Oct 2002 07:50:54 -0400
Subject: Source Code for Image Processing Functions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for a collection of highly optimized source code (in C) for the
basic image processing functions.

Everett Ramer
Cellomics, Inc.
Pittsburgh, PA


From daemon Thu Oct 17 08:12:54 2002



From: Pijpers,Frank :      fpijpers-at-nl.feico.com
Date: Thu, 17 Oct 2002 15:04:13 +0200
Subject: RE: Source Code for Image Processing Functions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


} -----Original Message-----
} From: Everett Ramer [mailto:eramer-at-cellomics.com]
} Subject: Source Code for Image Processing Functions

You may want to have a look at Intel's IPP (http://www.intel.com/software/products/ipp/ipp30/index.htm).

Sincerely,

Frank Pijpers
---
TEM Product Architect FEI
mailto:fpijpers-at-nl.feico.com


From daemon Thu Oct 17 09:05:47 2002



From: Mike Delannoy :      delannoy-at-jhmi.edu
Date: Thu, 17 Oct 2002 09:49:48 -0400
Subject: Re: Brain Fixative

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Karen,
Shrinkage or swelling of course is related to the osmolarity of the fix
(and vehicle). I like to run a
variety of fixes ie 2.5% GA vs 2% PF + 1.25% GA, etc. I think a good buffer
would be 0.1 M Hepes and
definitely add 3 mM CaCl2 for membrane tonicity. When osmicating it is
important to use KFECN6 (0.8%) reduced
osmium (1%), this keep the myelin sheaths from blebing (on ice). I also keep
the CaCl2 in the osmium step. I also
like long en-bloc staining (2% filtered UA for 2 hrs). This gives good
contrast and helps stabilize membranes during
dehydration. Remember even though you perfused to cut small pieces (3 mm
cubed) for infiltration of subsequnent solutions.
If you want to see actin or neurofilaments use 0.15% tannic acid after the
osmium (about 5 min only)-Kent McDonald-
With tissue I like to go 1:1 Propylene Oxide to Epon overnight. Good Luck.

Mike Delannoy
JHMI Microscopy Facility

"Jensen, Karen" wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Dear Listers:
}
} Does anyone have a recipe for mouse brain perfusion and fixation? I am
} trying to get the best brain preservation in mice. I trying to prevent
} neurons from swelling and I need to look at the Golgi apparatus. How about
} calcium chloride as an additive?
}
} Thanks!
}
} Karen
}
} Karen L. Bentley, M.S.(previously Jensen)
} Associate Scientist & Project Manager
} Electron Microscope Research Core
} University of Rochester Medical Center
} Rochester, NY 14642
} 585-275-1954



From daemon Thu Oct 17 09:17:10 2002



From: Ssjh1818-at-aol.com
Date: Thu, 17 Oct 2002 11:24:20 -0400
Subject: RE: Source Code for Image Processing Functions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


You might also check this page http://sourceforge.net/

Pavel

----- Original Message -----
} From: "Pijpers,Frank" {fpijpers-at-nl.feico.com}
To: "Microscopy-at-MSA.Microscopy.Com" {Microscopy-at-sparc5.microscopy.com}
Sent: Thursday, October 17, 2002 9:04 AM


we have an agfa duoscan T2500. we would like to better automate our scanning. does anyone know of a method of auto scaning negs on the duoscan.
thank s
john



From daemon Thu Oct 17 10:31:25 2002



From: Ssjh1818-at-aol.com
Date: Thu, 17 Oct 2002 11:22:08 -0400
Subject: Re: batch scanning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


we have an agfa duoscan T2500. we would like to better automate our scanning. does anyone know of a method of auto scaning negs on the duoscan.
thank s
john


From daemon Thu Oct 17 11:18:31 2002



From: Dusevich, Vladimir :      dusevichv-at-umkc.edu
Date: Thu, 17 Oct 2002 11:10:06 -0500
Subject: Antivibration platform

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

We are going to move our EM lab to a new place
(in the same building) and are planning to put an
anti-vibration platform under the column of our
FEG-ESEM XL30. Do we have to place a HV transformer
on the same platform (the transformer is right behind
the column)?

Any additional advise will be very helpful also,
especially experience with anti-vibration platforms
of different manufacturers.

Right now I can routinely make micrographs at magnifications
only up to 50,000. After hours, when vibrations are lower,
I can (sometimes) go up to 100,000-150,000. Can platform
improve performance of microscope in this range of magnifications?

Thank you,

Vladimir


Vladimir M. Dusevich, Ph.D.
Electron Microscope Lab Manager
3127 School of Dentistry
650 E. 25th Street
Kansas City, MO 64108-2784

Phone: (816) 235-2072
Fax: (816) 235-5524
Web: http://www.umkc.edu/dentistry/microscopy




From daemon Thu Oct 17 14:30:52 2002



From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Thu, 17 Oct 2002 15:20:24 -0400
Subject: Fwd: Re: batch scanning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



}
}
} FotoLook has to be run as an independent application (not from within
} Photoshop).
} Batches can be defined here.
} To repeat the same batch, drag the contents of the Done folder to the ToDO
} folder and run the batch again.
}
} At 11:22 AM 10/17/2002 -0400, you wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America

____________________________________________________________________________
Michael Cammer Analytical Imaging Facility Albert Einstein Coll. of Med.
Jack & Pearl Resnick Campus 1300 Morris Park Ave. Bronx, NY 10461
(718) 430-2890 Fax: 430-8996 URL: http://www.aecom.yu.edu/aif/




From daemon Thu Oct 17 15:32:18 2002



From: Rick Hugo :      hugo-at-pdx.edu
Date: Thu, 17 Oct 2002 13:24:03 -0700
Subject: Technics MIM-IV ion mill parts?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello all - anybody have spare parts for the Technics MIM-IV ion mill, or
know where I can find them? The anode tips in our ion guns (TSA 1-aperture)
are pretty hopelessly worn out. Any help would be greatly appreciated!

Rick

{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}
Rick Hugo, Ph.D.
Research Associate, Geomicrobiology and Electron Microscopy Lab
Department of Geology, Portland State University
Portland, OR 97207-0751
{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}



From daemon Thu Oct 17 15:45:46 2002



From: Kevin Frischmann :      kfrisch-at-amnh.org
Date: Thu, 17 Oct 2002 16:38:54 -0400
Subject: Re: Source Code for Image Processing Functions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


At 07:50 AM 10/17/02 -0400, Everett Ramer {eramer-at-cellomics.com} wrote:
} I am looking for a collection of highly optimized source code (in C) for the
} basic image processing functions.


Optimized for what?



**Please note new department name**

Kevin Frischmann, Laboratory Manager
Microscopy & Imaging Facility
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 USA

Phone: 212-313-7975
Fax: 212-496-3480
email: kfrisch-at-amnh.org



From daemon Thu Oct 17 16:01:28 2002



From: saram-at-duke.edu
Date: Thu, 17 Oct 2002 16:50:49 -0400 (EDT)
Subject: Re: en-block verses post-embedding staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Gang Ning is partially right.

For what it's worth:

UA is soluble in veranol acetate buffer at pH 4.5 and below.

However, UA is INSOLUBLE in cacodylate buffer at 4.5 and 3.5.

UA is soluble in cacodylate buffer at pH 0 and 0.5.

I don't know what happens between 0.5 and 3.5, but who wants to use a
buffer at that pH anyway? Somebody may have publised this info, but I
really don't need to know enough to chase it down. If one wanted to
stain at that low pH, cacodylate doesn't have much buffering capacity
down there.

Don't quote me on the exact pH's. I just tried this with pH paper and
tiny drops, not a pH meter.

I still recommend washing out the caco with the buffer that the UA is in
(or water) before using adding UA---that is if you choose to do en bloc
staining.

Sara Miller



On Tue, 15 Oct 2002, gang ning wrote:

} One thing I want to point out is people usually don't use UA in
} cacodylate buffer is NOT for the reason that UA is NOT SOLUBLE in it,
} but for the reason that UA works better in buffers of low pH.
}
} saram-at-duke.edu wrote:
}
} } Again, see CAPS below.
} } (CAPS ARE USED TO DIFFERENTIATE COMMENTS, NOT TO CONVEY HOSTILITY.)
} }
} }
} } On Sun, 13 Oct 2002, gang ning wrote:
} }
} } } Thanks for the clear definition of biohazard. But it is obvious that
} } } 1) UA is a nasty/hazard chemical. Everyone wants to avoid it if possible,
} } }
} }
} } UA CERTAINLY CAN BE CALLED NASTY. IT DOES REQUIRE SPECIAL HANDLING
} } PROCEDURES, AS DO OTHER CHEMICALS THAT WE USE AND CAN BE HANDLED SAFELY
} } UNDER THE PROPER CONDITIONS. NOT EVERYONE WANTS TO AVOID IT. FOLKS MUST
} } MAKE THEIR OWN JUDGEMENT AS TO WHETHER THE EXTRA PROCESSING TIME AND CARE
} } OF HANDLING ANOTHER CHEMICAL IS WORTH IT TO THEM. NOT USING UA IS FINE.
} } HOWEVER, DO NOT MAKE A GENERALIZATION ABOUT *EVERYONE* JUST BECAUSE YOU
} } CHOOSE NOT TO USE IT.
} }
} } } 2) it does take at least 1 hr + time for washing thoroughly (how long?)
} } } for a 1 mm3 block .... it is clearly time-consuming procedure
} } }
} }
} } IT DOES REQUIRE EXTRA TIME (1-2 HOURS). ONE CAN STAIN A 1 mm3 BLOCK FOR
} } 30-60 MIN AND WASH FOR 30-60 MIN.
} }
} } unless your turn around time is not an issue and working overtime is not a
} }
} } } matter with your hospital policy,
} } }
} }
} } TIME IS VERY MUCH AN ISSUE FOR US IN THE HOSPITAL SETTING. OUR TURN
} } AROUND TIME IS *ROUTINELY* NEXT DAY FOR SAMPLES FIXED IN GLUT, OS, UA.
} } TISSUE COMES IN IN THE AM, IS PROCESSED THAT DAY, BAKED OVERNIGHT, CUT AND
} } VIEWED THE NEXT DAY, AND PATHOLOGISTS HAVE MICROGRAPHS *IN HAND* THAT
} } AFTERNOON--ALMOST ALWAYS. MY TECHS RARELY HAVE TO WORK OVERTIME, AND WHEN
} } THEY DO, IT'S USUALLY BECAUSE OF EXTRA CUTTING OF HAVING TO RETRIEVE
} } SPECIMENS OUT OF PARAFFIN BLOCKS OR OFF SLIDES, NOT BECAUSE THEY PUT
} } SOMETHING INTO UA.
} }
} } YOU ARE CLEARLY VERY NEGATIVE TOWARD THE USE OF UA. FINE. DON'T USE IT.
} } BUT DON'T BE SARCASTIC ABOUT SPECIMEN TURN AROUND TIME AND
} } TECHNICIAN OVERTIME, BECAUSE IT JUST ISN'T AN ISSUE IF SPECIMENS AND
} } WORKLOADS ARE MANAGED PROPERLY.
} }
} } } 3) en block staining has adversary effect on showing glycogen so is not
} } } suitable for muscle, heart, liver and so on; there is not significant
} } } meaning for kidney .. then what else left? LOTS:
} } }
} }
} } BRAIN, CILIA, TUMORS, VIRUSES, MUSCLE DISEASES OTHER THAN GLYCOGEN STORAGE
} } DISEASES, RESEARCH SAMPLES, ETC.
} }
} }
} } It is overkill for solid tissue blocks, IT IS NOT OVERKILL WHEN THE
} } BEST PRESERVATION, RATHER THAN DECREASING PROCESSING TIME IS OF
} } IMPORTANCE.
} }
} } and 4) ethanolic/methanolic UA of higher concentration penetrates tissue
} }
} } } much faster than other aqueous UA solutions and gives fresher image and
} } } better contrast but sometimes leave precipitation around cell member no
} } } matter how one washes before staining.
} } }
} }
} } AS I SAID BEFORE, IF THE CORRECT BUFFER IS USED SO THAT UA DOES NOT
} } PRECIPTATE WITH IT, THERE IS NO PRECIPITATION. PERHAPS YOUR *HIGHER
} } CONCENTRATION* IS A CAUSE OF DIFFICULTY IN WASHING. SINCE UA IS SOLUBLE
} } IN ALCOHOL, UNBOUND, UNPRECIPITATED UA IS EASILY WASHED OUT BY THE
} } DEHYDRATING AGENT(S). IF YOU HAVE PRECIPITATE, IT IS MOST LIKELY DUE TO
} } SOME REACTION WITH THE SOLUTION LEFT IN THE CELL BEFORE YOU ADD THE UA, OR
} } PERHAPS THE *HIGHER* CONCENTRATION YOU USE ISN'T WASHED OUT???
} }
} } } For EM work, the simpler and better, as long as it works.
} } }
} }
} } YES, THE KEY IS, *AS LONG AS IT WORKS* FOR THE PURPOSE YOU HAVE. IF THE
} } PURPOSE IS TO PRODUCE THE BEST ULTRASTRUCTURE, AND IF MEMBRANES AND OR
} } LIPIDS ARE PART OF THE QUESTION, USE OF UA IS GOOD.
} }
} } Overkill is always a problem. YES, NO ONE WANTS TO SPEND MORE TIME THAN
} } IS NECESSARY OR USE CHEMICALS THAT CAN BE HAZARDOUS IF NOT HANDLED
} } PROPERLY, BUT UA IS NOT ALWAYS OVERKILL.
} }
} } IT DEPENDS ON YOUR APPLICATION. EACH TO HIS OWN.
} }
} } }
} } }
} } } saram-at-duke.edu {mailto:saram-at-duke.edu} wrote:
} } }
} } } } See CAPS interspersed below for distinction from the original.
} } } }
} } } } Sara Miller
} } } }
} } } } On Fri, 11 Oct 2002, Gang Ning wrote:
} } } }
} } } } } ------------------------------------------------------------------------
} } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com {mailto:ListServer-at-MSA.Microscopy.Com}
} } } } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } } -----------------------------------------------------------------------.
} } } } }
} } } } }
} } } } } Hi
} } } } }
} } } } } I agree with Dr. Sara Miller's point that UA en block staining in some extent
} } } } } helps in preservation of ultrastructure as well as enhancement of the
} } } } } contrast. However, it also has disadvantages such as one has to deal with UA,
} } } } } a biohazard reagent,
} } } } }
} } } } A BIOHAZARD IS A BIOLOGICAL AGENT THAT IS HAZARDOUS (INFECTIOUS). UA IS
} } } } RADIOACTIVE AND IS A HEAVY METAL. IT, LIKE GLUTARALDEHYDE, OSMIUM, LEAD,
} } } } RESINS, PROPYLENE OXIDE, ETC., SHOULD BE HANDLED WITH CARE. HANDLED
} } } } PROPERLY, IT IS NOT ANY MORE DANGEROUS THAN THE OTHER NASTY CHEMICALS
} } } } ELECTRON MICROSCOPISTS HAVE TO USE. IT IS NOT A BIOLOGICAL AGENT.
} } } }
} } } } and sometimes background from crystal precipitation,
} } } }
} } } } } especially when ethanolic UA is used.
} } } } }
} } } } WE NEVER SEE PRECIPITATION. PERHAPS TRY DILUTING THE STAIN OR WASHING
} } } } THOROUGHLY BEFORE USING IT WILL ELIMINATE THE PRECIPITATE. UA IS NOT
} } } } SOLUBLE IN PHOSPHATE OR CACODYLATE BUFFER; IF YOU DON'T WASH OUT THESE
} } } } BUFFERS BEFORE ADDING UA IN VERANAL ACETATE BUFFER, UA IN WATER, OR UA IN
} } } } ETHANOL OR METHANOL, YOU CAN GET PRECIPATE.
} } } }
} } } } More importantly, it is time-consuming
} } } }
} } } } } because it takes hours for aqueous UA solution to penetrate a sizable tissue
} } } } } block.
} } } } }
} } } } WE USE 1 MM TISSUE BLOCKS AND STAIN FOR AN HOUR 1% UA IN 0.11M VERONAL
} } } } ACETATE BUFFER. IF THE TISSUE OR CELL PELLET IS SMALLER, YOU CAN GET AWAY
} } } } WITH 30 MIN IN UA.
} } } }
} } } } Therefore, I would say it all depends your tissue and purpose of your
} } } }
} } } } } analysis.
} } } } }
} } } } I AGREE. IT ALL DEPENDS ON YOUR PURPOSE. IF YOU WANT BEAUTIFUL
} } } } ULTRASTRUCTURE AND YOU'RE NOT LOOKING SPECIFICALLY FOR GLYCOGEN, USE UA EN
} } } } BLOCK. IF YOU'RE IN A HURRY OR YOU CARE TO SEE ALL THE GLYGOGEN, DON'T
} } } } USE IT.
} } } }
} } } } } In my lab, we routinely do en block with samples of blood, tissue culture,
} } } } } virus, bacteria and other microorganisms, which are believed more vulnerable
} } } } } to solvent, but not with solid tissue. We have a lot of samples from kidney
} } } } } biopsy and we didn't find en block made any difference from staining thin
} } } } } sections whereas taking your time to have a sufficient osmication is much
} } } } } more helpful to your structure.
} } } } }
} } } } } Greg Ning
} } } } }
} } } } }
} } } } }
} } } } } mohammed y abdulrawoof / f40z006 75760 wrote:
} } } } }
} } } } } } ------------------------------------------------------------------------
} } } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } } } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com {mailto:ListServer-at-MSA.Microscopy.Com}
} } } } } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } } } } -----------------------------------------------------------------------.
} } } } } }
} } } } } } Dear listers,
} } } } } } Please advise as to the usefulness/advantage of doing enblock staining of
} } } } } } biological tissues as against doing heavy-metal staining of grids with
} } } } } } thin sections. The tissues that I am about to process are for diagnosis
} } } } } } (pathology).
} } } } } }
} } } } } } Thanks
} } } } } } M. Yousuf Abdul-Rawoof
} } } } } }
} } } } } --
} } } } } Gang Ning, M.D., Ph.D.
} } } } } Director
} } } } } Electron Microscopy Facility
} } } } } Medical College of Wisconsin
} } } } } 8701 Watertown Plank Road
} } } } } Milwaukee, Wisconsin 53226
} } } } } (414) 456-8344
} } } } } e-mail: gning-at-mcw.edu {mailto:gning-at-mcw.edu}
} } } } } http://www.mcw.edu/cellbio/emfacility.html
} } } } }
} } } } }
} } } } }
} } } } }
} } } } }
} } } } Sara E. Miller, Ph. D.
} } } } P. O. Box 3712
} } } } Duke University Medical Center
} } } } Durham, NC 27710
} } } } Ph: 919 684-3452
} } } } FAX: 919 684-3265
} } } }
} } } }
} } } --
} } }
} } } Gang Ning, M.D., Ph.D.
} } }
} } } Director
} } }
} } } Electron Microscopy Facility
} } }
} } } Medical College of Wisconsin
} } }
} } } 8701 Watertown Plank Road
} } }
} } } Milwaukee, WI 53226
} } }
} } } (414) 456-8141
} } }
} } }
} } }
} } }
} }
} } Sara E. Miller, Ph. D.
} } P. O. Box 3712
} } Duke University Medical Center
} } Durham, NC 27710
} } Ph: 919 684-3452
} } FAX: 919 684-3265
} }
} }
}
} --
}
} Gang Ning, M.D., Ph.D.
}
} Director
}
} Electron Microscopy Facility
}
} Medical College of Wisconsin
}
} 8701 Watertown Plank Road
}
} Milwaukee, WI 53226
}
} (414) 456-8141
}
}
}

Sara E. Miller, Ph. D.
P. O. Box 3712
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-3265



From daemon Fri Oct 18 02:52:41 2002



From: pvosta-at-unionbio-eu.com
Date: Fri, 18 Oct 2002 09:40:32 +0200
Subject: Re: Source Code for Image Processing Functions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I have put together an extensive collection of links to all kinds of
microscopy and imaging resources on the WWW and there are several links
to software and source code repositories:

http://ourworld.compuserve.com/homepages/pvosta/pcrimg.htm

Best regards,

Peter Van Osta

==============================================
Everett Ramer wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} I am looking for a collection of highly optimized source code (in C) for the
} basic image processing functions.
}
} Everett Ramer
} Cellomics, Inc.
} Pittsburgh, PA


From daemon Fri Oct 18 05:27:56 2002



From: Kevin Mackenzie :      k.s.mackenzie-at-abdn.ac.uk
Date: Fri, 18 Oct 2002 11:10:31 +0100
Subject: 30th SCOTTISH MICROSCOPY SYMPOSIUM Wednesday 13th November

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Final Announcement

30th SCOTTISH MICROSCOPY SYMPOSIUM Wednesday 13th November 2002

Aberdeen Exhibition & Conference Centre, Bridge of Don, Aberdeen AB23 8BL

Registration Form available on web site
http://www.abdn.ac.uk/emunit/smg2002.htm

Programme

10.00-10.45 "Mechanical Interactions of Cells with Surfaces" by Mathis
Riehle (Glasgow University)

10.45-11.30 Applications of Microscopical Techniques in Forensic Science"
by Alison Crossley, (AEA Technology plc, Oxford)

11.30-11.45 "FTIR Microscopy of Human Trabecular Bone" by Alison Coates,
(University of Aberdeen)

11.45-12.00 "Techniques for visualisation of apoptosis in osteoclasts" by
Mike Rogers and F. Coxon (University of Aberdeen Medical School)

12.00-13.45 Luncheon, Trade Exhibition, Poster Display

13.45-14.30 "Creating, Viewing and Storing Large Image Data Fields" by Alan
Entwistle, (Ludwig Institute, University of London)

14.30-14.45 "Development of an Image Analysis Program for Measuring Bone
Resorption in vitro" by Rob van't Hof and Miep Helfrich (University of
Aberdeen Medical School)

14.45-15.00 "Microscopy of Tick Salivary Glands" Barry Stewart (Zoology,
University of Aberdeen)

15.00-15.30 TEA

15.30-16.15 "Microscopy and variant Creutzfeldt-Jakob (vCJD) disease" by
James Ironside, (CJD Surveillance Unit, Edinburgh)

16.15-16.30 'Iso-butanol/water: a new staining technique for light and
electron microscopy by Ian Roberts', (Scottish Crop Research Institute, Dundee)

16.30 Presentation of Prizes



Electron Microscope unit
School of Biological Sciences
University of Aberdeen
Aberdeen
AB24 2TZ

Tel 01224-272847
Fax 01224-272396
------------
Kevin Mackenzie
k.s.mackenzie-at-abdn.ac.uk



From daemon Fri Oct 18 09:44:52 2002



From: Fred Pearson :      eoptics-at-mcmail.cis.mcmaster.ca
Date: Fri, 18 Oct 2002 10:31:46 -0400 (EDT)
Subject: Electropolishing Solution required

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Dear Group:

I have the task of preparing TEM foils from 1 mm. O.D. 18K and 14K Gold
wire. I was successful in ion milling dimpled 50um section/foils in PIPS
at 4Kv., 4 degree angle. We are looking at the strain fields about 100nm.
precipitates. The ion milling caused damaged to the structure as indicated
by the ripple patterns in the material as viewed in the TEM. Obviously
this may/can alter our examination.

Next, I would like to try electropolishing the material if I can make
suitable foils for the polisher. We cold rolled the wire, and punched out
3 mm. disks to work with. Electropolishing is the next step to try.

What I am asking, does anyone know of an electropolishing solution for Au,
suitable for a Tenupol 5 or a Fischione machine?

Thanks in advance

Fred Pearson


********************************************************
Fred Pearson
Brockhouse Institute for Materials Research
McMaster University
1280 Main St. West
Hamilton, Ontario
Canada L8S 4M1


email: eoptics-at-mcmaster.ca
phone: (905) 525-9140 ext. 24609
fax: (905) 521-2773
********************************************************



From daemon Fri Oct 18 10:38:11 2002



From: saram-at-duke.edu
Date: Fri, 18 Oct 2002 11:27:00 -0400 (EDT)
Subject: Re: Prions and OsO4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Phil, et al.

Thanks for your comments, Phil.

Listers, I'm forwarding this answer to the list because of our
recent discussion on UA and the question posed by Phil about UA
inactivation of prions.

The question at a recent microscopy meeting concerned whether osmium
"killed" (inactivated) prions. It arose because of a question about how
specimens suspected of containing prion material (Creutzfeld-Jacob
Disease and other spongiform encephalapathies) should be handled in the EM
lab.

The discussion centered around osmium inactivation because it is a proven
fact that aldehydes DO NOT inactivate prions. Osmium is usually the next
step, and most EM folks use it; some do not use UA. The osmium
inactivation question was also visited on this net some years ago.

My answer was that some oxidizing agents (peracetic acid; Vet J 159:10,
2000) reduce activity of prions in brain, and thus, one might suspect
that osmium could act similarly. This article does say that many
oxidizing agents tried have no effect, and that the effect is related to
whether the prions are aggretated (the more aggregated, as in homogenized
specimens, the less inactivated). Peracetic acid works better on brain
than on homogenized tissue. However, I pointed out that in a personal
communication with both Drs. D. M. Taylor and Stanley Prusiner, these
experts simply said that the experiment of osmium inactivation hasn't been
done. We simply don't know. Also, they pointed out that we don't even
know whether binding up the prions in resin inactivates them. They
recommended working with gloves and collecting the "chips" from trimming
for incineration.

Phil has raised the question of whether UA would inactivate prions. His
reasoning has merrit. However, I would maintain that, like osmium, UA has
not been tested, and that some things that one would think would
inactivate prions don't (see ref above). Just because they are stained or
bound up doesn't necessarily mean they are inactive.

Radioactivity, even very strong doses, does not inactivate prions. As a
matter of fact, this is one reason why folks feel that prions don't
contain nucleic acids.

If histology labs have to handle brain suspected of containing prions,
they inactivate the tissue with 97% formic acid after formaldehyde
fixation (Neurology 40:887, 1990), which doesn't seem too detrimental to
structure.


Sara Miller


On Fri, 18 Oct 2002, Philip Oshel wrote:

} Sara,
}
} I've been mulling over your "does OsO4 oxidize prions?" question that
} you mentioned in your talk at the meeting here in Madison this week.
} Why OsO4? It's a poor protein stain, so although it's a strong
} oxidizer, it wouldn't occur to me to suspect it of degrading prions.
} Why not uranyl acetate or (better?) uranyl nitrate? These are good
} protein stains, and so should bind better to prions. Further,
} although their radioactivity is low, it is present, and as alpha
} particles. Nice and heavy, and they do lots of damage at short range.
} I should think UAc or UNO3 would be useful for breaking down prions.
} (I refuse to say "kill", as I don't think prions have any of the
} attributes of life.)
}
} Mind, I'm not going to do the experiment either. 8-)
}
} Nice talk!
}
} Phil
} --
} Philip Oshel
} Supervisor, BBPIC microscopy facility
} Department of Animal Sciences
} University of Wisconsin
} 1675 Observatory Drive
} Madison, WI 53706 - 1284
} voice: (608) 263-4162
} fax: (608) 262-5157 (dept. fax)
}
}
}

Sara E. Miller, Ph. D.
P. O. Box 3712
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-3265



From daemon Fri Oct 18 12:30:15 2002



From: Goodhouse, Joseph :      jgoodhouse-at-molbio.Princeton.EDU
Date: Fri, 18 Oct 2002 13:15:53 -0400
Subject: Tracking software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I need to hear back from those who have said they have software for tracking
cells and particle movement. Please let me know if you are ready to bring
in demo version.

Regards,

Joe Goodhouse
Confocal / EM Core Laboratory
Department of Molecular Biology
Princeton University
609-258-5432

Visit us at http://www.molbio.princeton.edu/facility/confocal/index.html



From daemon Fri Oct 18 13:52:27 2002



From: Richard Edelmann :      edelmare-at-muohio.edu
Date: Fri, 18 Oct 2002 14:42:45 -0400
Subject: RE: WORM drives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


(Oh, ye who scoff at SCSI in favor of IDE)

Richard:

Easy:

Adaptec SCSI + EasySCSI + Win4.0 = it just works.

Win NT directly supports MO drives. I just went and pulled my old LF-
7303 off the shelf, checked the SCSI ID wasn't in conflict, plugged it into a
"Newly rebuilt" NT box, rebooted, and Ta-Da there it was. The SCSI interface
itself handles everything.

Panasonic actually has a BIOS update for it:

http://www.panasonic.com/support/software/pd_other.htm



http://www.panasonic.com/support/software/pd_other.htm


} If your drive is the old WORM technology, good luck finding one! I haven't seen
} one for sale in years.
}
} What computer does the 360 use? I have forgotten... Even if it is a DEC
} (perhaps especially) it would be easier to send the data to a newer PC for
} storage on hard drives and CD-Rs.
}
} Regards, Woody
}
} Woody White
} McDermott Technology Inc.
} McD: http://www.mtiresearch.com/
} Mine: http://woody.white.home.att.net
}
}
} -----Original Message-----
} } From: Richard Beanland [mailto:richard.beanland-at-bookham.com]
} Sent: Wednesday, October 16, 2002 9:12 AM
} To: 'Microscopy-at-MSA.Microscopy.Com'
} Subject: WORM drives
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hello everyone,
} although this is strictly not a microscopy question, I'm posting
} this in the hope that someone has been through this process already and can
} offer some advice.
}
} Our LEO 360FE SEM computer saves images on a Panasonic LF-7300 WORM drive.
} It doesn't seem feasible to connect the computer to our intranet directly
} (too old/difficult!) but a relatively straighforward way to get at the
} images is to install an identical WORM drive on a PC which is networked.
} However I imagine that some kind of software is needed to make the drive
} functional under Windows NT. My internet searches for drivers have drawn
} blanks so far. Has anyone done this and can give me some advice?
} We are asking LEO for some support on this but independent advice is
} always very valuable.
}
} Many thanks,
}
} Richard
}
}
} _______________________________
} Richard Beanland
} Analytical Services
} Bookham Technology plc
} Caswell,
} Towcester,
} Northamptonshire NN12 8EQ
} UK
} Tel: +44 (0) 1327 356362
} Fax: +44 (0) 1327 356398
} http://www.bookham.com
}
}
}
} =======================================================================
} This e-mail is intended for the person it is addressed to only. The
} information contained in it may be confidential and/or protected by
} law. If you are not the intended recipient of this message, you must
} not make any use of this information, or copy or show it to any
} person. Please contact us immediately to tell us that you have
} received this e-mail, and return the original to us. Any use,
} forwarding, printing or copying of this message is strictly prohibited.
}
} No part of this message can be considered a request for goods or
} services.
} =======================================================================
} Any questions about Bookham's E-Mail service should be directed to
} postmaster-at-bookham.com.
}
}
}



Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Supervisor
350 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513.529.5712 Fax: 513.529.4243
E-mail: edelmare-at-muohio.edu
http://www.emf.muohio.edu

"RAM disk is NOT an installation procedure."


From daemon Fri Oct 18 14:31:45 2002



From: Peter Tomic :      PTomic-at-anadigics.com
Date: Fri, 18 Oct 2002 15:30:53 -0400
Subject: Foil covered SEM stubs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


FOLKS;

Our analytical chemist stated that aluminum foil has a coating of a
lubricant on it to facilitate separation from the roll. I was wondering if
this has caused anyone any issues with EDX.

Peter Tomic
Anadigics, Inc.

-----Original Message-----
} From: Tina Carvalho [mailto:tina-at-pbrc.hawaii.edu]
Sent: Wednesday, October 16, 2002 1:54 PM
To: Microscopy Listserver


Hi, all-

Here is a colleague's method for using aluminum foil on SEM stubs for a
smooth background:

1. "Reynolds" seems smoothest and "regular" seems easiest to work with.

2. Burnishing. I've tried various things and they usually end up
scratching the foil. To make sure it is absolutely smooth flat, I
carefully cut a strip that is a bit wider and longer than a stub.
Place it on a clean surface, cover it with something very smooth
(weighing paper works well) and work over it with a finger tip to get
it absolutely as flat as possible. Picking it up by the very edge
with tweezers helps avoid wrinkles.

3. The smooth, flat piece is then transferred to a piece of
double-stick scotch tape and once more pressed to the tape to get a
very flat surface with no bubbles between foil and tape. This should
be done on glass or a surface from which tape plus foil can then be
carefully lifted.

4. The tape plus foil can then be applied to the stub - if it still
looks nice, proceed to step five. If wrinkles or bumps have appeared,
pull it up and try again. (You can put the tape directly on the stub
and then apply the foil - However, I have found that it works best
for me to have a nicely bonded duo of smooth tape and foil first)

5. Using weighing paper, again, press tape plus foil gently to the stub.

6. Finally, use a single-edge razor blade to trim off the excess foil
and tape around the edge of the stub. You can apply a couple of teeny
dabs of silver paste at several edge points if you want to make
absolutely certain the foil is well connected to the stub. I have
gotten into the habit of making the strip of foil plus tape slightly
narrower than stub diameter so that there is a small crescent of stub
above and below the strip, simply as a way of telling top and bottom
at a glance.

7. Additional note. For mounting an object with a convex (like my
little larval shells) or irregular lower surface, aluminum foil can
be gouged with a minuten insect pin before applying silver paste or
paint to help achieve better contact and bonding.

As with everything, it just takes a bit of experimentation. . . .


I hope this helps!

Aloha,
Tina



From daemon Fri Oct 18 14:46:58 2002



From: Mike Coviello :      coviello-at-mae.uta.edu
Date: Wed, 09 Oct 2002 14:40:39 -0700
Subject: EM-Looking for a full time electronics service engineer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi All:
We are looking for a highly qualified, skilled individual to service
several TEM's, SEM's and other related electronic equipment across our
campus. The position is full time and permanent. The qualified
individual should be able to troubleshoot and repair many brands of
electronic equipment down to the component level.
If interested, please forward a resume to me.
Thanks,
Michael Coviello
Lab Manager,
The University of Texas at Arlington,
Arlington, TX



From daemon Fri Oct 18 16:20:26 2002



From: Lawrence Mason :      LMason-at-wyeth.com
Date: Fri, 18 Oct 2002 17:05:36 -0400
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe



From daemon Fri Oct 18 17:06:33 2002



From: JoAn Hudson :      hudson-at-uoneuro.uoregon.edu
Date: Fri, 18 Oct 2002 15:02:32 -0700
Subject: Looking for book reviewers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I have 4 books ready for review. If you would like to review any of
these titles and write your review for the journal please let me know
ASAP. Reviewers will be chosen on a first come first serve basis based
on your interests. I would also like to know if you would review other
books as they come available if your selection has already been sent
out.

List:

1) Electron Backscatter Diffraction in Materials Science, Edited by
Adam J. Schwartz, Mukul Kumar, and Brent L. Adams

2) Scanning Tunneling Microscopy and its Applications, By C.Bai

3) PCR/RT-PCR in situ, light and electron microscopy, By Gerard
Morel and Mireille Raccurt

4) Quick Photoshop for research, a guide to digital imaging for
photoshop 4x,5x,6x,7x, by Jerry Sedgewick

If interested please e-mail me along with a mailing address for you.
Thank you very much!
JoAn

JoAn Hudson, PhD.
Institute of Neuroscience
222 Huestis Hall
University of Oregon,
Eugene, OR 97403
 
Telephone: (541) 346 4508
FAX: (541) 346 4548






From daemon Fri Oct 18 18:01:55 2002



From: David Henriks :      henriks-at-southbaytech.com
Date: Fri, 18 Oct 2002 15:53:45 -0700
Subject: Electropolishing Solution required

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Fred:

I will apologize in advance in that my recipe is for our Model 550D
Single Vertical Jet ElectroPolisher rather than the other systems you
mentioned. Seems like there is always a solution with one of our
products! ;-)

In any case, I do have a paper from Bernie Kestel titled:

"Jet Thinning of YBa2Cu3Ox High Tc Superconductor and Gold for TEM with
a Non-Acid Electrolyte"

This is Technical Report #14 on our online list. Unfortunately, it is
not in electronic format, but the reference is Ultramicroscopy 25 (1988)
351-354. I can also send you a photocopy of the paper if that would be
easier for you.

While some of our recipes appear to be interchangeable with the twin jet
systems. I believe that his BK-2 solution (non-acid electrolyte) is too
viscous to be pumped through the twin jet units. However, it may give
you a good starting point.

DISCLAIMER: South Bay Technology, Inc. produces equipment and supplies
as described above and, therefore, has a vested interest in promoting
their use.

Best regards-

David

Fred Pearson wrote:


------------------------------------------------------------------------

The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

-----------------------------------------------------------------------.

Dear Group:

I have the task of preparing TEM foils from 1 mm. O.D. 18K and 14K
Gold
wire. I was successful in ion milling dimpled 50um section/foils in
PIPS
at 4Kv., 4 degree angle. We are looking at the strain fields about
100nm.
precipitates. The ion milling caused damaged to the structure as
indicated
by the ripple patterns in the material as viewed in the TEM. Obviously

this may/can alter our examination.

Next, I would like to try electropolishing the material if I can make
suitable foils for the polisher. We cold rolled the wire, and punched
out
3 mm. disks to work with. Electropolishing is the next step to try.

What I am asking, does anyone know of an electropolishing solution for
Au,
suitable for a Tenupol 5 or a Fischione machine?

Thanks in advance

Fred Pearson

********************************************************
Fred Pearson
Brockhouse Institute for Materials Research
McMaster University
1280 Main St. West
Hamilton, Ontario
Canada L8S 4M1

email: eoptics-at-mcmaster.ca
phone: (905) 525-9140 ext. 24609
fax: (905) 521-2773
********************************************************

--
David Henriks
Vice President

South Bay Technology, Inc.
1120 Via Callejon
San Clemente, CA 92673 USA

TEL: +1-949-492-2600
Toll-free in the USA: +1-800-728-2233
FAX: +1-949-492-1499

email: henriks-at-southbaytech.com

Celebrating 38 years of providing Materials Processing Solutions for
Metallogaphy, Crystallography and Electron Microscopy.

Please visit us online at www.southbaytech.com.

The information contained in this message and any attachments is
privileged and confidential. This message is intended for the
individual or entity addressed.
If you are not the intended recipient, please do not read, copy or
disclose this communication. Notify the sender of the mistake by
calling +1-949-492-2600 and
delete this message from your system.




From daemon Fri Oct 18 21:14:04 2002



From: Greg Barclay :      gbarclay-at-trinidad.net
Date: Fri, 18 Oct 2002 22:01:04 -0500
Subject: Thanks for the Stereo/Compound LM replies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to eveyone who vetured to reply to my only half tongue in
cheek posting about which microscopes are really compound. I
learned that some stereomicroscopes have only one objective lens
(!?) which must call some interesting optical designs.




From daemon Fri Oct 18 21:28:51 2002



From: Greg Barclay :      gbarclay-at-trinidad.net
Date: Fri, 18 Oct 2002 22:18:02 -0500
Subject: Tissue shrinkage in Tousimis CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I am seeking advice on the correct way to operate a Tousimis semiautomatic CPD to minimize if not eliminate
plant tissue shrinkage.

I would appreciate any suggestions.

Thank you.

Greg Barclay




From daemon Sat Oct 19 06:44:48 2002



From: dieter-at-magnet.at (Dieter Aigner)
Date: Sat, 19 Oct 2002 13:30:05 +0200
Subject: Need help with LEITZ 1512 microtome

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sirs,

I have received an used Leitz 1512 microtome a few days ago and are not
very
experienced with it. Everything works perfectly but I can not manage it to
set the
THICKNESS to MORE THAN 11 my. I can rotate the knob only from 0-11.
Is the microtome broken?
I do not have a manual for the microtome.

Please can anyone help me?

Many thanks.

Dieter, AUSTRIA




From daemon Sun Oct 20 14:46:33 2002



From: Diane G. Miller :      millerd-at-coho.net (by way of MicroscopyListserver)
Date: Sun, 20 Oct 2002 14:13:45 -0500
Subject: Convert bitmap to tiff or jpeg

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello All,

I was wondering if some of you could forward me some sites for converting
bitmap images to tiff or jpeg. I would appreciate it very much.

Thanks for your help.
Diane

Diane G. Miller
Miller Consultant Service
503-784-6444
millerd-at-coho.net


From daemon Sun Oct 20 20:05:38 2002



From: Philip Oshel :      peoshel-at-facstaff.wisc.edu
Date: Sun, 20 Oct 2002 19:55:16 -0500
Subject: Re: Tissue shrinkage in Tousimis CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Greg,

This depends a great deal on the nature and size of the plant
tissues. We do a lot of small, herbaceous specimens here, mostly
Arabidopsis, but I've done other plant tissues as well, such as roots.
The trick is to ignore the directions in the Tousimis manual -- for
any biological specimens. What you need to do instead is to let the
specimens soak in the liquid CO2 for a given amount of time after
first purging the chamber of the absEtOH that was in it when the
samples were loaded. Then purge the chamber to replace the lq CO2
with fresh lq CO2, soak, and repeat N times. This is analogous to the
dehydration process where the water in the tissues is replaced with
EtOH. Now, the EtOH in the tissues is being replaced with lq CO2.
Once that is done, then and only then can the samples be dried.
Otherwise, they are being dried from EtOH.
The length of the soaks and number of soaks is rather empirical and
highly dependent on tissue type and size. For the small (a few mm),
herbaceous Arabidopsis specimens, we generally use 4 soaks for 10
minutes each. For 1 cm^2 bone, I've used 5 soaks of 30 - 60 minutes
each. For fetal rat urogential sinus, 3 soaks of 2 to 3 minutes each
works. For small (meaning a cm to mm) crustaceans and their
appendages, which have relatively impermeable cuticles, I used 5
soaks of 5 or 10 minutes each.
There can be some extraction of materials by the lq CO2 and possibly
some artifacts from too long in lq CO2, but it is a much worse sin to
run the specimens through CO2's critical point while they still
retain EtOH. It is imperative to make certain that *all* of the EtOH
has been replaced by lq CO2.
This means you'll have to try different times and number of soaks.
The larger and denser or woodier your samples, the more soaks for
longer times you'll need.
Some of the times I've given could perhaps be shortened. Given that
too long likely won't produce significant artifact (note the
"likely") and will only take more time, but that too little time will
certainly produce ruined specimens which wastes a *lot* more time,
specimens, and more, I'd rather spend the (possibly) extra time on
the EtOH/lq CO2 exchanges.

Phil

} I am seeking advice on the correct way to operate a Tousimis
} semiautomatic CPD to minimize if not eliminate
} plant tissue shrinkage.
}
} I would appreciate any suggestions.
}
} Thank you.
}
} Greg Barclay

--
}}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
Philip Oshel
Supervisor, AMFSC and BBPIC microscopy facilities
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706 - 1284
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)


From daemon Mon Oct 21 00:03:59 2002



From: Thomas Litzinger :      thomas_c_litzinger-at-yahoo.com
Date: Sun, 20 Oct 2002 21:54:08 -0700 (PDT)
Subject: Re: Tissue shrinkage in Tousimis CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe


__________________________________________________
Do you Yahoo!?
Y! Web Hosting - Let the expert host your web site
http://webhosting.yahoo.com/


From daemon Mon Oct 21 04:57:00 2002



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Mon, 21 Oct 2002 10:46:55 +0100 (GMT Daylight Time)
Subject: Re: Tissue shrinkage in Tousimis CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


You said re CPD:

"There can be some extraction of materials by the lq CO2
and possibly some artifacts from too long in lq CO2...."

We are doing a student project here into SEM preparation
methods for pollen. I wonder if you could let me know if
you have any references for the above comment?

Thanks

Dave


On Sun, 20 Oct 2002 19:55:16 -0500 Philip Oshel
{peoshel-at-facstaff.wisc.edu} wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Greg,
}
} This depends a great deal on the nature and size of the plant
} tissues. We do a lot of small, herbaceous specimens here, mostly
} Arabidopsis, but I've done other plant tissues as well, such as roots.
} The trick is to ignore the directions in the Tousimis manual -- for
} any biological specimens. What you need to do instead is to let the
} specimens soak in the liquid CO2 for a given amount of time after
} first purging the chamber of the absEtOH that was in it when the
} samples were loaded. Then purge the chamber to replace the lq CO2
} with fresh lq CO2, soak, and repeat N times. This is analogous to the
} dehydration process where the water in the tissues is replaced with
} EtOH. Now, the EtOH in the tissues is being replaced with lq CO2.
} Once that is done, then and only then can the samples be dried.
} Otherwise, they are being dried from EtOH.
} The length of the soaks and number of soaks is rather empirical and
} highly dependent on tissue type and size. For the small (a few mm),
} herbaceous Arabidopsis specimens, we generally use 4 soaks for 10
} minutes each. For 1 cm^2 bone, I've used 5 soaks of 30 - 60 minutes
} each. For fetal rat urogential sinus, 3 soaks of 2 to 3 minutes each
} works. For small (meaning a cm to mm) crustaceans and their
} appendages, which have relatively impermeable cuticles, I used 5
} soaks of 5 or 10 minutes each.
} There can be some extraction of materials by the lq CO2 and possibly
} some artifacts from too long in lq CO2, but it is a much worse sin to
} run the specimens through CO2's critical point while they still
} retain EtOH. It is imperative to make certain that *all* of the EtOH
} has been replaced by lq CO2.
} This means you'll have to try different times and number of soaks.
} The larger and denser or woodier your samples, the more soaks for
} longer times you'll need.
} Some of the times I've given could perhaps be shortened. Given that
} too long likely won't produce significant artifact (note the
} "likely") and will only take more time, but that too little time will
} certainly produce ruined specimens which wastes a *lot* more time,
} specimens, and more, I'd rather spend the (possibly) extra time on
} the EtOH/lq CO2 exchanges.
}
} Phil
}
} } I am seeking advice on the correct way to operate a Tousimis
} } semiautomatic CPD to minimize if not eliminate
} } plant tissue shrinkage.
} }
} } I would appreciate any suggestions.
} }
} } Thank you.
} }
} } Greg Barclay
}
} --
} }}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
} Philip Oshel
} Supervisor, AMFSC and BBPIC microscopy facilities
} Department of Animal Sciences
} University of Wisconsin
} 1675 Observatory Drive
} Madison, WI 53706 - 1284
} voice: (608) 263-4162
} fax: (608) 262-5157 (dept. fax)
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



From daemon Mon Oct 21 05:14:37 2002



From: Richard Beanland :      richard.beanland-at-bookham.com
Date: Mon, 21 Oct 2002 11:06:30 +0100
Subject: WORM drives - thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I had several very useful replies, thankyou to everyone who took the time to
help.
Although I am now able to connect and have the drive working on a WinNT
PC, it appears that the disks are unreadable. The LEO 'scope has an
operating system which pre-dates Win95 (even Win3.1, I think!), and the file
allocation tables used on the MO disk can't be deciphered by the modern PC.
So, it looks like I'll have to explore other routes. At the moment this
looks like getting a CD-RW drive (if possible, I suspect there will be
problems along this route too). I feel it should be possible to put a cable
between adjacent boxes of electronics and send signals from one to another
without the need for a physical medium inbetween, but maybe not...

Thanks again,

Richard

_______________________________
Richard Beanland
Analytical Services
Bookham Technology plc
Caswell,
Towcester,
Northamptonshire NN12 8EQ
UK
Tel: +44 (0) 1327 356362
Fax: +44 (0) 1327 356398
http://www.bookham.com



=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.

No part of this message can be considered a request for goods or
services.
=======================================================================
Any questions about Bookham's E-Mail service should be directed to postmaster-at-bookham.com.



From daemon Mon Oct 21 07:12:45 2002



From: White, Woody N. :      nwwhite-at-mcdermott.com
Date: Mon, 21 Oct 2002 06:59:05 -0500
Subject: RE: Convert bitmap to tiff or jpeg

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Most image editors will let you save images in a multitude of formats. I
your case it would be load-resave steps. Some may allow batch processing.

Examples:
IrView http://www.irfanview.com/
Or VuePrint http://www.hamrick.com/

Regards, Woody




Woody White
McDermott Technology Inc.
McD: http://www.mtiresearch.com/
Mine: http://woody.white.home.att.net






------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hello All,

I was wondering if some of you could forward me some sites for converting
bitmap images to tiff or jpeg. I would appreciate it very much.

Thanks for your help.
Diane

Diane G. Miller
Miller Consultant Service
503-784-6444
millerd-at-coho.net


From daemon Mon Oct 21 08:49:44 2002



From: Gilles Hug :      gilles.hug-at-onera.fr
Date: Mon, 21 Oct 2002 15:40:11 +0200
Subject: Re: Antivibration platform

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Hi,

Le jeudi, 17 oct 2002, à 18:10 Europe/Paris, Dusevich, Vladimir a écrit
:

}
} We are going to move our EM lab to a new place
} (in the same building) and are planning to put an
} anti-vibration platform under the column of our
} FEG-ESEM XL30. Do we have to place a HV transformer
} on the same platform (the transformer is right behind
} the column)?

Yes if the transformer is linked to the microscope with a rigid cable
it would be betterto have it on the platform. Unless it generates by
itself mechanical vibrations.

}
} Any additional advise will be very helpful also,
} especially experience with anti-vibration platforms
} of different manufacturers.

We have recently built a platform for a new microscope. It is composed
of a concrete block. I am sending you also pdfs of recent measurements.
As you can see on the graphs we are rather satisfied of the results.
The filtering system consists of a concrete bloc of 15 Mkg (tons)
suspended with 4 springs with dampers at each corners. The company Gerb
delivered the system. They made a calculation that the cutoff frequency
would be around 2-3 Hz. Generally speaking such arrangement can be view
as to spring in series (on for the block and one for the microscope).
You must take care that there is no resonance between the to
oscillators. In my case I made a quick estimate, which showed that this
is not a problem. In fact as a rule of thumb the weight of the
microscope should be negligible as compared to the platform. So design
you platform with 10 times the weight of the microscope.
The shape is also important. The vertical motion is normally the more
important mode. But, if for example the shape of the concrete block is
tall rather than flat, you may be annoyed by rotations around X, Y
axis. This might excite flexion modes of the column, especially if
their is some attachments like X-ray detector..
In our case we also have a preexisting concrete floor that is probably
very thick but unfortunately connected to the building (old building).
This probably why we can see the vibrations coming from the air
conditioning which is located on the roof. If you make a new building
you should consider building a concrete floor *de-connected* to the
building and the your platform on dampers.
Also, I have to mention that the operator is NOT on the platform but on
a wooden floor above it.
Finally, last advice and not the least: check EVERYTHING YOURSELF.
Generally, the workers don't understand why you ask specific things and
if they don't take care they can ruin your beautiful design. We had a
mason who accidentally poured liquid concrete that you use to flatten
the floors. Some of it went below the platform and re-connected it with
the building!!! Fortunately if was not very difficult to remove. But
what would happen if we haven't noticed?

Note about the graphs: please note that the vertical scales are
correct. These measurements have been performed by our "vibration”
department at Onera. Usually , they study vibrations modes of Aibuses
and other flying machines. They had to build a special low noise
amplifier for those measurements. This is the reason why we can see
these very small contributions. I guess that with commercial equipments
you would not be able to see these peaks above the noise.
Best,
Gilles

PDFs : ask me if interested by the figures


From daemon Mon Oct 21 09:57:36 2002



From: Jensen, Karen :      Karen_Jensen-at-urmc.rochester.edu
Date: Mon, 21 Oct 2002 10:47:01 -0400
Subject: Renal EM Job

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers:

Our renal pathologist is moving to Columbus, Ohio and is taking a position
at Ohio State. He is looking for a trained EM tech who is experienced in
renal pathology EM. Contact me off list directly to my email address if
anyone is interested.

karen_bentley-at-urmc.rochester.edu

Karen


Karen L. Bentley, M.S.(previously Jensen)
Associate Scientist & Project Manager
Electron Microscope Research Core
University of Rochester Medical Center
Rochester, NY 14642
585-275-1954



From daemon Mon Oct 21 10:00:55 2002



From: Ian MacLaren :      maclaren-at-tu-darmstadt.de
Date: Mon, 21 Oct 2002 16:54:03 +0200
Subject: Alternatives to C films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,
What are the alternatives to C-support films, for supporting powder
particles. We have a situation where we want to determine the C content
of an inorganic phase and don't want a C film, which would mess up the
quantification.

Thanks
--
Ian MacLaren
Strukturforschung, Materialwissenschaft, TU-Darmstadt, Germany
ian.maclaren-at-physics.org / http://homepages.tu-darmstadt.de/~maclaren/



From daemon Mon Oct 21 10:10:28 2002



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Mon, 21 Oct 2002 10:03:37 -0500
Subject: Re: WORM drives - thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello again. Sorry to hear about the difficulties.

The first order of business will be to find out what system is running the
LEO. After that is done, you can probably explore options. They may not all
be that much fun. There is a possibility that there will not only be an
operating system incompatibility, but also a file format compatibility.

We were running a Kevex EDS system when the Internet started coming on big.
We much wanted to make files available over the net - me especially because
I had my office in another building. The Kevex operated around a DEC PDP-11
computer with the TSX+ operating system. (Those aren't exactly common.) The
first order of business was to connect that to the ethernet. We were able
to do it but it required a special card costing several hundred dollars
(now you can buy $10 ethernet cards for PCs) and TCP-IP software costing
many hundreds more (compared to the built in support in modern OSs). Then
we had to convert the proprietary Kevex file format over to a portable
format (TIF). Once we had crossed those two not-so-small hurdles, we were
up and running just fine.

The other approach, like you have been trying, was to find a common media
format so that we could transfer removable media between two computers.
Someone had written a utility to read PDP-11 disks on a PC (getting around
the different allocation and directory structures), but I think it may have
been limited to floppy disks. We were using 10 MB Bernoulli cartridges from
Iomega. I vaguely recall that there were other issues trying to access
those disks. Even if we had succeeded, there would still have been the
issue of converting the file formats.

I would like to think that it will not be so bad for you. I would like to
think that the LEO is built around a PC. But LEO or another LEO user would
probably have to verify that. They should also be able to tell which OS and
what disk system are being used. Then you could move on from there.

Warren

At 11:06 AM 10/21/02 +0100, you wrote:

} I had several very useful replies, thankyou to everyone who took the time to
} help.
} Although I am now able to connect and have the drive working on a WinNT
} PC, it appears that the disks are unreadable. The LEO 'scope has an
} operating system which pre-dates Win95 (even Win3.1, I think!), and the file
} allocation tables used on the MO disk can't be deciphered by the modern PC.
} So, it looks like I'll have to explore other routes. At the moment this
} looks like getting a CD-RW drive (if possible, I suspect there will be
} problems along this route too). I feel it should be possible to put a cable
} between adjacent boxes of electronics and send signals from one to another
} without the need for a physical medium inbetween, but maybe not...
}
} Thanks again,
}
} Richard

-------------------------------------------
No files should be attached to this message
-------------------------------------------
Warren E. Straszheim, Ph.D.
Materials Analysis and Research Lab
Iowa State University
23 Town Engineering
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
Web: www.marl.iastate.edu

Scanning electron microscopy, x-ray analysis, and image analysis of materials
Computer applications and networking




From daemon Mon Oct 21 10:26:56 2002



From: Dmitry Lobanov :      dima-at-ipm.sci-nnov.ru (by way of
Date: Mon, 21 Oct 2002 10:06:30 -0500
Subject: TEM Need help on Ion Etching Device

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Now I am looking for an electrical motors

5116-207GE M:1516 E 0065 + G:15/3 (reduction ratio 900:1)
and
5116-211GE M:1516 E 0065 + G:15/3 (reduction ratio 1670:1)

They are need for positioning and rotation of specimen table in
BALZERS Universal Ion Beam Etching Device IEU 100.

With kind regards,
Dmitri Lobanov.


Institute for physics of microstructures RAS
603950, N.Novgorod, GSP-105, Russia
phone: +7(8 312)675037
fax: +7(8 312)609111
E-mail: dima-at-ipm.sci-nnov.ru


From daemon Mon Oct 21 10:30:36 2002



From: David_Bell-at-millipore.com
Date: Mon, 21 Oct 2002 11:24:08 -0400
Subject: Tensile Stage in ESEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello fellow microscopists,

I am looking for a lab that has a tensile stage mounted in an ESEM that is
willing to run a couple of samples. I would prefer a service lab, however
if there is a commercial or university lab that is willing to take on the
task, that would be fine as well. The samples would be polymer films and
I would be interested in images corresponding to the various force data.
If you have this capability, please contact me off list.

Thanks in advance,



David Bell
Scientist
Electron Microscopy Lab
Millipore Corporation
80 Ashby Road
Bedford, MA 01730
(781) 533-2108
David_Bell-at-Millipore.com


From daemon Mon Oct 21 10:54:40 2002



From: Berg, R. Howard :      RHBerg-at-danforthcenter.org
Date: Mon, 21 Oct 2002 10:48:56 -0500
Subject: variable temperature hot plate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,

Does anyone know a source for a variable temperature hot plate? We
want to use it for wax embedding, for pouring in situ boats, allowing
us to slide em over and readjust the tissue as it solidifies.


Howard


From daemon Mon Oct 21 10:58:48 2002



From: Jensen, Karen :      Karen_Jensen-at-urmc.rochester.edu
Date: Mon, 21 Oct 2002 11:51:44 -0400
Subject: Diagnostic EM Job

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers:

Is anyone interested in a diagnostic EM job? You must have experience in
renal pathology and tumors. Our renal pathologist is moving to Ohio and is
looking for anyone in the USA who is interested and available.

Please contact me off list at: karen_bentley-at-urmc.rochester.edu

Karen L. Bentley, M.S.(previously Jensen)
Associate Scientist & Project Manager
Electron Microscope Research Core
University of Rochester Medical Center
Rochester, NY 14642
585-275-1954



From daemon Mon Oct 21 13:19:12 2002



From: Walck, Scott D. :      walck-at-ppg.com
Date: Mon, 21 Oct 2002 14:09:49 -0400
Subject: RE: Convert bitmap to tiff or jpeg

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


My favorite is ThumbsPlus
http://www.cerious.com/

-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8515 (fax)



-----Original Message-----
} From: White, Woody N. [mailto:nwwhite-at-mcdermott.com]
Sent: Monday, October 21, 2002 7:59 AM
To: 'Diane G. Miller'; microscopy-at-sparc5.microscopy.com


Most image editors will let you save images in a multitude of formats. I
your case it would be load-resave steps. Some may allow batch processing.

Examples:
IrView http://www.irfanview.com/
Or VuePrint http://www.hamrick.com/

Regards, Woody




Woody White
McDermott Technology Inc.
McD: http://www.mtiresearch.com/
Mine: http://woody.white.home.att.net






------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hello All,

I was wondering if some of you could forward me some sites for converting
bitmap images to tiff or jpeg. I would appreciate it very much.

Thanks for your help.
Diane

Diane G. Miller
Miller Consultant Service
503-784-6444
millerd-at-coho.net


From daemon Mon Oct 21 13:36:54 2002



From: Shu-You Li :      syli-at-northwestern.edu
Date: Mon, 21 Oct 2002 13:28:11 -0500
Subject: Immediate Postdoc position available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
I am posting this message for Prof. Dravid. Please contact him directly.
Thank you,
Shuyou.
=======================================================
IMMEDIATE POSTDOCTORAL POSITION AVAILABLE:

CRYO-ANALYTICAL TRANSMISSION ELECTRON MICROSCOPY
OF SOFT AND HYBRID NANOSTRUCTURES

INSTITUTE FOR ENVIRONMENTAL CATALYSIS, NORTHWESTERN UNIVERSITY

Applications are invited for a full-time postdoctoral research associate within the Institute for Environmental Catalysis (IEC) at Northwestern University. The position is jointly supervised by Profs. Vinayak P. Dravid of Materials Science & Engineering, and Prof. Jean-Francois Gaillard of Civil Engineering, at Northwestern University.

We seek a dynamic and energetic individual with prior experience in advanced electron microscopy of micobiological, geochemical, or materials systems to conduct frontier research using state-of-the-art TEM instrumentation; and design novel approaches to observe aquatic environmental particles in-situ in TEM. Hands-on experience in preparatory methods such as cryo-preparation, ultramicrotomy etc. will be a strong asset, as well as knowledge of STEM-EDS, PEELS, and energy filtering.

Applications should include a vita, a brief statement of research interests, and names of at least three references including email addresses and telephone numbers.

The position is open only to those eligible for employment in the US.

Applications in PDF format should be sent by email to: v-dravid-at-northwestern.edu

Professor Vinayak P. Dravid,
Materials Science & Engineering, and
Institute for Environmental Catalysis
2225 N. Campus Drive, 1133 Cook Hall (formerly MLSF)
Northwestern University, Evanston, IL 60208, USA
Ph.: (847) 467-1363, Fax: (847) 467-6573
E-mail: v-dravid-at-northwestern.edu
http://vpd.ms.northwestern.edu
http://epic.ms.northwestern.edu

_____________________________
Shu-You Li, Ph.D.
Electron Microscopist
Electron Probe Instrumentation Center(EPIC)
Northwestern University
2220 Campus Drive, 1141 Cook Hall
Evanston, IL 60208, USA
Ph: (847) 491-7798, Fax: (847) 491-7820
Email: syli-at-northwestern.edu; syli16-at-hotmail.com
http://pubweb.northwestern.edu/~sli974





From daemon Mon Oct 21 14:05:49 2002



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Mon, 21 Oct 2002 14:57:44 -0500
Subject: Non-carbon-containing TEM support films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Ian MacLaren wrote:
=========================================================
What are the alternatives to C-support films, for supporting powder
particles. We have a situation where we want to determine the C content of
an inorganic phase and don't want a C film, which would mess up the
quantification.
==========================================================
There are two possibilities:

a) silicon dioxide/monoxide films made via the evaporation of SiO, see URL
http://www.2spi.com/catalog/spec_prep/evapor_3b.shtml or the completed
films can be purchased already made from SPI Supplies (and some of our
competitors), see URL
http://www.2spi.com/catalog/grids/cusctgrd.html

or

b) silicon nitride membrane window TEM grids as shown on URL
http://www.2spi.com/catalog/instruments/silicon-nitride.shtml

Either way, you will have a good support that will be completely free of any
carbon.

Disclaimer: SPI Supplies provides all three items mentioned, namely the SiO
chips, already custom coated SiO2 TEM grids, and the silicon nitride
membrane window grids.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================




From daemon Mon Oct 21 14:17:07 2002



From: Becky Holdford :      r-holdford-at-ti.com
Date: Mon, 21 Oct 2002 14:10:04 -0500
Subject: contract micro-FTIR work in Texas area?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


If anyone has any information about labs that do contract
micro-FTIR work in the Texas area, please forward it to me.
We are looking for a source for this type of work closer
than Sunnyvale, CA.

--
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Becky Holdford (r-holdford-at-ti.com)
972-995-2360
972-648-8743 (pager)
SC Packaging FA Development
Texas Instruments, Inc.
Dallas, TX
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~




From daemon Mon Oct 21 14:35:51 2002



From: Lucille A. Giannuzzi :      lag-at-mail.ucf.edu
Date: Mon, 21 Oct 2002 15:23:04 -0400
Subject: TEM Specimen Preparation Course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


A TEM Specimen Preparation Course Will be offered March 12-14, 2003 at the
Materials Characterization Facility at the University of Central Florida,
Orlando, FL USA

A review of all techniques will be given, with primary emphasis on tripod
polishing, ion milling, and all latest FIB techniques

Instructors:
Ron Anderson, Microscopy Today
Fred Stevie, NC State
Lucille Giannuzzi, UCF
Brian Kempshall, UCF

For more information contact Lucille Giannuzzi, lag-at-mail.ucf.edu

*******************************************************************
Lucille A. Giannuzzi, Ph.D.

Associate Professor, Mechanical, Materials, and Aerospace Engineering
Director, UCF/Cirent Materials Characterization Facility
University of Central Florida
12443 Research Parkway, Suite 305
Orlando, FL 32826 email lag-at-mail.ucf.edu phone (407) 882-1500
fax (407) 275-4321
--------------------------------------------------------------------
"Good judgement comes from experience.

Experience comes from making bad judgement."

Mark Twain
********************************************************************




From daemon Mon Oct 21 14:35:53 2002



From: pkrsko-at-stevens-tech.edu (by way of MicroscopyListserver)
Date: Mon, 21 Oct 2002 14:24:37 -0500
Subject: Ask-A-Microscopist:Calibration Standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (pkrsko-at-stevens-tech.edu) from
http://www.msa.microscopy.com/Ask-A-Microscopist.html on Monday,
October 21, 2002 at 13:31:19
---------------------------------------------------------------------------

Email: pkrsko-at-stevens-tech.edu
Name: Peter Krsko

Organization: Stevens Institute of Technology

Education: Graduate College

Location: Hoboken, NJ 07030

Question: Dear Sirs,

Where can I purchase some kind of calibration sample for fluorescent
imaging? I would like to qauntify my results, but I am not sure what
intensity corresponds to what concentration of FITC. Are there also
calibration samples for reflective fluoro microscopy?

Thank you very much

---------------------------------------------------------------------------


From daemon Mon Oct 21 14:39:51 2002



From: Lucille A. Giannuzzi :      lag-at-mail.ucf.edu
Date: Mon, 21 Oct 2002 15:31:43 -0400
Subject: FIB FIG CALL FOR PAPERS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


CALL FOR PAPERS FIB Focused Interest Group (FIG) WORKSHOP

Tuesday, March 18, 2003

Student Union, University of Central Florida, Orlando, FL 32816

(in conjunction with a TEM Specimen Prep Course March 12-14, 2003, and the
FL AVS/Florida Society for Microscopy Meeting, March 17-18, 2003).

Please submit an abstract (limit 250 words), including authors and
affiliations, by email to:

Dr. Lucille Giannuzzi
lag-at-mail.ucf.edu

*******************************************************************
Lucille A. Giannuzzi, Ph.D.

Associate Professor, Mechanical, Materials, and Aerospace Engineering
Director, UCF/Cirent Materials Characterization Facility
University of Central Florida
12443 Research Parkway, Suite 305
Orlando, FL 32826 email lag-at-mail.ucf.edu phone (407) 882-1500
fax (407) 275-4321
--------------------------------------------------------------------
"Good judgement comes from experience.

Experience comes from making bad judgement."

Mark Twain
********************************************************************




From daemon Mon Oct 21 16:04:30 2002



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Mon, 21 Oct 2002 16:48:30 -0600
Subject: Muscle TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear List,

I would appreciate recieveing any fixation protocol you know
works well for mouse muscle (extensor digitorum longus (EDL)
and epitrochlearus (EPI)). We use routinely immersion into
2.5% glutaraldehyde in 0.1M cacodylate pH 7.4, but have had
often problems with mitochondrial swelling. Since this project is
to quantify mitochondrial density, we do need perfect and
reliable fixation of all mitos. Perfusion is not an option. Thanks
for any advise on type of fixative, buffer and temperature.

Marc
--
Marc Pypaert
Department of Cell Biology,
Center for Cell and Molecular Imaging and
Ludwig Institute for Cancer Research,
Yale University School of Medicine
333 Cedar Street
New Haven CT 06520
Tel : (203) 785 3681
Fax : (203) 785 7446




From daemon Mon Oct 21 17:21:48 2002



From: Walck, Scott D. :      walck-at-ppg.com
Date: Mon, 21 Oct 2002 18:09:25 -0400
Subject: Need a high temperature epoxy for microtoming

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Why don't you embed the samples in glass? You are at the softening temperatures of glass. We have a technique here that we bond two pieces of glass together by heating them in a plasma etching system. This technique was originally developed and published by Prof Carlos Pantano at Penn State University to profile Sn on the bottom surface of glass. What I would do is sprinkle some of your dust on one of the glass surfaces, bring the two glass surfaces together and heat them up. Then dimple and ion mill normally. I have no idea whether you particles will wet the glass or not, but they might.

Another option is to use a ceramic bonding adhesive in a similar manner to above. There is a description of a high temperature XTEM prep in the MRS TEM sample Prep book number 1 on what to use. The adhesive was from Aremco. Take two piece of Si and put your adhesive on one surface, dust the adhesive and make your sandwich. Alternatively, you could just embed them in the ceramic adhesive and prepare the disk as a plan view sample.

If you choose one of these ideas, I would love to know how it works out.

-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8515 (fax)

"The opinions expressed are those of Scott D. Walck and not of PPG Industries, Inc. nor of any PPG-associated companies."




-----Original Message-----
} From: "George_Munzing-at-engelhard.com"-at-sparc5.microscopy.com
[mailto:"George_Munzing-at-engelhard.com"-at-sparc5.microscopy.com]
Sent: Tuesday, October 15, 2002 9:23 AM
To: Microscopy-at-sparc5.microscopy.com


Hello fellow microscopists,

I need to do some high temperature TEM studies and am looking for a
suitable mounting resin. I will be embedding and microtoming powder
materials and would like to go as high in temperature as possible. The
materials themselves can withstand the temperatures but I doubt my
conventional epoxy would survive.

Is there any silicone epoxy or other suitable material designed to
withstand high temperatures (maybe as high as 600-800C) for a short time
for embedding?

Any suggestions you can offer are greatly appreciated.

Thanks in advance!

George R. Munzing Jr.
Engelhard Corporation
Strategic Technologies Group
25 Middlesex/Essex Turnpike
Iselin, NJ 08830
Tel: 732-205-7030
FAX: 732-205-5300



From daemon Mon Oct 21 17:43:09 2002



From: Arrowood, Roy :      arrowood-at-utep.edu
Date: Mon, 21 Oct 2002 16:36:10 -0600
Subject: variable hot plate source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Howard,

Cole-Parmer lists temperature-controlled hot plates (made by Heisen, and
probably others)--from fixed setpoint units with +/-1 degree temperature
control to fancy programmables. I'm pretty sure that the other major lab
supply houses (Fisher, VWR, etc.) also offer variable-temperature hot
plates, but I don't have catalogs handy at the moment, so I can't verify
that.

You might also consider using a Variac(TM) (or other variable
autotransformer) to power a basic, old-fashioned clunker of a hot plate. If
you use that approach, make sure the power capacity of the autotransfomer
matches the power of the hot plate. Don't buy a Variac that would let users
dial in too high a voltage, which could fry the hot plate and/or present
safety hazards. And don't use the transformer approach if your hot plate
has fancy extras such as stirmotors, digital displays, etc.---many devices
other than plain-vanilla electrical resistance heaters can be damaged by
undervoltage. Also, the autotransformer strategy would require that you
invest some time to "calibrate" your setup: i.e,, to find out at what
voltage the hot plate maintains a satisfactory and safe temperature for
what you want to do. Changing the environment or setup (anything that
increases or decreases heat losses from the hot plate) would Furthermore,
autotransformers are pricey items (as are T-controlled hot plates).

So your best bet, probably, is to buy a nice temperature-controlled hot
plate, as you indicated.

Hope this helps.

-------Roy

P.S. I have no connection to any of the suppliers or manufacturers
mentioned, and no preference for any supplier is intended. Shop with care
to make sure anything you buy meets your requirements.


======================
Roy Arrowood arrowood-at-utep.edu
Metallurgical and Materials Engineering
University of Texas at El Paso
El Paso, TX 79968-0520
(915)747-6934 voice // (915)747-8036 FAX
(915)747-5468 secretary



From daemon Mon Oct 21 23:39:20 2002



From: Tim Richardson :      mirlyn-at-attglobal.net
Date: Mon, 21 Oct 2002 18:40:15 -0600
Subject: Synchronization of cell cultures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a method or protocol for synchronizing cells in tissue
culture?

We are looking to find a method so that all cells in the culture are at
substantially the same point in their cell cycle? It is also important
that the cells are still alive and living as normally as cells do when in
tissue culture.

We are most interested in achieving this goal with human fibroblasts in
culture, but would be interested in any success with other cells lines of
any origin, mammal, animal, or plant.

Thank you for your assistance, any information on achieving this goal will
be helpful.

Tim Richardson,
RTI,
Phone: 905-951-7058
Fax: 905-951-7052

email: mirlyn-at-attglobal.net





From daemon Tue Oct 22 02:31:48 2002



From: William.Ho-at-glencore.com.sg
Date: Tue, 22 Oct 2002 15:15:01 +0800
Subject: pls take me off the e-mail list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


i have no idea why i'm on it in the first place



From daemon Tue Oct 22 06:51:25 2002



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Mon, 21 Oct 2002 10:46:55 +0100 (GMT Daylight Time)
Subject: Re: Tissue shrinkage in Tousimis CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


You said re CPD:

"There can be some extraction of materials by the lq CO2
and possibly some artifacts from too long in lq CO2...."

We are doing a student project here into SEM preparation
methods for pollen. I wonder if you could let me know if
you have any references for the above comment?

Thanks

Dave


On Sun, 20 Oct 2002 19:55:16 -0500 Philip Oshel
{peoshel-at-facstaff.wisc.edu} wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Greg,
}
} This depends a great deal on the nature and size of the plant
} tissues. We do a lot of small, herbaceous specimens here, mostly
} Arabidopsis, but I've done other plant tissues as well, such as roots.
} The trick is to ignore the directions in the Tousimis manual -- for
} any biological specimens. What you need to do instead is to let the
} specimens soak in the liquid CO2 for a given amount of time after
} first purging the chamber of the absEtOH that was in it when the
} samples were loaded. Then purge the chamber to replace the lq CO2
} with fresh lq CO2, soak, and repeat N times. This is analogous to the
} dehydration process where the water in the tissues is replaced with
} EtOH. Now, the EtOH in the tissues is being replaced with lq CO2.
} Once that is done, then and only then can the samples be dried.
} Otherwise, they are being dried from EtOH.
} The length of the soaks and number of soaks is rather empirical and
} highly dependent on tissue type and size. For the small (a few mm),
} herbaceous Arabidopsis specimens, we generally use 4 soaks for 10
} minutes each. For 1 cm^2 bone, I've used 5 soaks of 30 - 60 minutes
} each. For fetal rat urogential sinus, 3 soaks of 2 to 3 minutes each
} works. For small (meaning a cm to mm) crustaceans and their
} appendages, which have relatively impermeable cuticles, I used 5
} soaks of 5 or 10 minutes each.
} There can be some extraction of materials by the lq CO2 and possibly
} some artifacts from too long in lq CO2, but it is a much worse sin to
} run the specimens through CO2's critical point while they still
} retain EtOH. It is imperative to make certain that *all* of the EtOH
} has been replaced by lq CO2.
} This means you'll have to try different times and number of soaks.
} The larger and denser or woodier your samples, the more soaks for
} longer times you'll need.
} Some of the times I've given could perhaps be shortened. Given that
} too long likely won't produce significant artifact (note the
} "likely") and will only take more time, but that too little time will
} certainly produce ruined specimens which wastes a *lot* more time,
} specimens, and more, I'd rather spend the (possibly) extra time on
} the EtOH/lq CO2 exchanges.
}
} Phil
}
} } I am seeking advice on the correct way to operate a Tousimis
} } semiautomatic CPD to minimize if not eliminate
} } plant tissue shrinkage.
} }
} } I would appreciate any suggestions.
} }
} } Thank you.
} }
} } Greg Barclay
}
} --
} }}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{
} Philip Oshel
} Supervisor, AMFSC and BBPIC microscopy facilities
} Department of Animal Sciences
} University of Wisconsin
} 1675 Observatory Drive
} Madison, WI 53706 - 1284
} voice: (608) 263-4162
} fax: (608) 262-5157 (dept. fax)
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



From daemon Tue Oct 22 07:15:01 2002



From: Mary Gail Engle :      mgengle-at-pop.uky.edu
Date: Tue, 22 Oct 2002 08:07:19 -0400
Subject: Re: Muscle TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Marc,
We use to routinely fix mouse muscle by immersion but used 3.5% glut. in
.1M cacodylate at 4 degrees for 1.5 hours. The pieces need to be very
small. Add 4% sucrose to the buffer washes. The pH should be 7.4. Osmium
should be 1% in the same buffer at 4 degrees also.
Good luck,
Mary Gail

At 04:48 PM 10/21/02 -0600, Marc Pypaert wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Mary Gail Engle
Sr. Research Laboratory Manager
Electron Microscopy & Imaging Facility
Health Sciences Research Bldg. 001
University of Kentucky
Lexington, KY 40536-0305

phone 859-323-6108
fax 859-257-9700


From daemon Tue Oct 22 09:07:50 2002



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Tue, 22 Oct 2002 09:48:46 -0400
Subject: Re: Muscle TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Marc,
I've always had goo results in muscle (skeletal and cardiac) with
the following fix:
4% pfa (methanol-free, diluted from 16% stock )
2.5% glutaraldehyde
in 0.1M Na-cacodylate buffer
My "recipe", making this up from 10ml stock vials of 16% pfa and 10%
glut yields 40 ml of fix, to which I then add 2 ml of saturated
picric acid (aqueous). (based on Ito & karnovsky, J Cell Bio,
abstract 168a, 1968)

Try to hold the muscles in an extended state, either by leaving them
attached to the leg bone, or using ligature to tie them to small
wooden sticks. This will help prevent overly contracted sarcomeres
and give a more "textbook" appearance with M-lines, A & I bands, etc.
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From daemon Tue Oct 22 09:10:52 2002



From: David BG Rose :      drose-at-wlgore.com
Date: Tue, 22 Oct 2002 10:05:11 -0400
Subject: EM service lab in Taiwan

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello Listers,

An associate of mine is looking for a service lab in Taiwan that would be
able to pot and polish samples and examine the samples with a SEM. Any
information as to company names and/or contacts would be appreciated.

Thank you.

- David






From daemon Tue Oct 22 10:37:07 2002



From: saram-at-duke.edu
Date: Tue, 22 Oct 2002 11:25:20 -0400 (EDT)
Subject: Re: Muscle TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


There is at least one study showing that the important thing, as far a
shrinkage or swelling is concerned, is the osmolarity of the buffer. This
should be close to physiological (300 mOsm). Glutaraldehyde will add
about 100 mOsm for each percent, thus, a 1% glutaraldehyde solution in
buffer that would be 300 mOsm by itself would be ~400 mOsm. However, the
tonicity added by the fixaive apparently doesn't detract from the
morphology.

The 0.1 M cacodylate + 4% sucrose mentioned below will be close to the
percent we used to use with good success. I think, however, our final
sucrose concentration was 3.8%. We used to make up double strength buffer
and double strength aqueous fix and add them together 1:1; e.g., 0.2 M
cacodylate containing 6.8% sucrose and 4 % aqueous glut. Our final
concentration after mixing them 1:1 was 0.1 M caco, 3.4% sucr, 2% glut.
I think 3.5% glut is slight overkill, but won't hurt anything. Some
folks use 5%. You could also make the buffer with an increased
concentration of cacodylate, but it is expensive. 1M is sufficient
buffering capacity, and the sucrose makes up the rest of the tonicity.

Cacodylate will give a finer grained ultrastructure than phosphate, but
some folks don't want to use the arsenical. If you choose phosphate,
0.135 M phosphate buffer is physiological; you can also use less phosphate
and make up the difference with cheap sucrose. If you use something that
is ionic, like NaCl, remember that it dissociates into 2 ions and you have
to use half as much. Also, the 3.4 % sucrose translates in molarity to
0.1 M, thus if you use an ionic substitute, use molarity to calculate the
amount, not percent. A 1M solution of sucrose and a 0.5 M solution of
NaCl will give the same osmolarity.

Additionally, a small amount (2.5 mM, i.e., 0.0025 M) of calcium is good
for membranes. This is not enough to alter the osmolarity very much.

The other thing you need to do if you're interested in muscle bands (and
maybe mitochondrial size???) is to fasten the muscle bundle in a
muscle clamp before immersing in fix. Otherwise, it contracts.

This may be more than you wanted to know, but hope some of it helps.

Sara Miller


On Tue, 22 Oct 2002, Mary Gail Engle wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Marc,
} We use to routinely fix mouse muscle by immersion but used 3.5% glut. in
} .1M cacodylate at 4 degrees for 1.5 hours. The pieces need to be very
} small. Add 4% sucrose to the buffer washes. The pH should be 7.4. Osmium
} should be 1% in the same buffer at 4 degrees also.
} Good luck,
} Mary Gail
}
} At 04:48 PM 10/21/02 -0600, Marc Pypaert wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } Dear List,
} }
} } I would appreciate recieveing any fixation protocol you know
} } works well for mouse muscle (extensor digitorum longus (EDL)
} } and epitrochlearus (EPI)). We use routinely immersion into
} } 2.5% glutaraldehyde in 0.1M cacodylate pH 7.4, but have had
} } often problems with mitochondrial swelling. Since this project is
} } to quantify mitochondrial density, we do need perfect and
} } reliable fixation of all mitos. Perfusion is not an option. Thanks
} } for any advise on type of fixative, buffer and temperature.
} }
} } Marc
} } --
} } Marc Pypaert
} } Department of Cell Biology,
} } Center for Cell and Molecular Imaging and
} } Ludwig Institute for Cancer Research,
} } Yale University School of Medicine
} } 333 Cedar Street
} } New Haven CT 06520
} } Tel : (203) 785 3681
} } Fax : (203) 785 7446
} }
} }
}
}
} Mary Gail Engle
} Sr. Research Laboratory Manager
} Electron Microscopy & Imaging Facility
} Health Sciences Research Bldg. 001
} University of Kentucky
} Lexington, KY 40536-0305
}
} phone 859-323-6108
} fax 859-257-9700
}
}
}

Sara E. Miller, Ph. D.
P. O. Box 3712
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-3265



From daemon Tue Oct 22 11:00:39 2002



From: mohammed y abdulrawoof / f40z006 75760 :      mdyousuf-at-KSU.EDU.SA
Date: Tue, 22 Oct 2002 19:44:18 -0300 (GMT)
Subject: Used UA and OsO4 - Disposal