From: Nestor J. Zaluzec :      zaluzec-at-microscopy.com
Date: Sat, 1 May 2004 08:58:52 -0500
Subject: [Microscopy] Administrivia: April Archives now on-line

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Colleagues

The April Archives are now on-line.

http://www.microscopy.com/MicroscopyListserver


Nestor
Your Friendly Neighborhood SysOop


From MicroscopyL-request-at-ns.microscopy.com Sat May 1 08:59:30 2004



From: esem1-at-traceevidence.org (by way of MicroscopyListserver)
Date: Sat, 1 May 2004 08:59:54 -0500
Subject: [Microscopy] viaWWW: Looking For Used SEM, EDS and Coating Equipment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (esem1-at-traceevidence.org) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Saturday, May 1, 2004 at 08:36:12
---------------------------------------------------------------------------

Email: esem1-at-traceevidence.org
Name: David Spence

Organization: Southwestern Institute of Forensic Sciences

Title-Subject: [Microscopy] [Filtered] MListserver:Looking For Used SEM, EDS and Coating Equipment

Question: I am looking for used SEM, EDS, sputter/carbon coater and related equipment. If you have or know of any equipment that is serviceable and available for haul-off or very inexpensive, please let me know. I am specifically looking for an older quality SEM with a motorized stage that is suitable for automated gunshot residue analysis.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sun May 2 13:14:45 2004



From: Steve Beck :      becks-at-sunynassau.edu
Date: Sun, 02 May 2004 14:17:59 -0400
Subject: [Microscopy] 3-D Imaging & Projection

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Colleagues,

I am involved with a consortium that is being partially funded by the
NSF-ATE program. The main project involves the development of 3-D
images related to various science, math, engineering and technology
areas (from molecules to organs to CAD/architecture) so that they can
be projected (stereographic projection). Each institution
participating in the project will have a 3-D theatre built so that
SMET classes can be brought in for demonstrations of course specific
images (the theatre will house a Barco 3-D projector driven by a SGI
UNIX-based workstation and a special screen that retains
polarization). Students and faculty from the partner institutions
will also be trained in a residential program at Brookhaven National
Lab by their 3-D Visualization Team.

Since I have been teaching TEM and SEM courses at NCC for almost 20
years, I thought that it would be an interesting project to render
student and my own SEM & TEM images for 3-D projection. I have seen
many threads on 3-D imaging on this listserver and I would appreciate
some advice on where to begin. Initially, I am supposed to identify
some software applications that would be suitable for rendering 3-D
images of my micrographs. I am aware of some of the applications that
are used at BNL for their 3-D theatre such as OpenDX, Open GL and
Visualization ToolKit (VTK).

Can anyone suggest an application (UNIX based ideally) that I might
get started with? As a novice in this area I realize that the concept
of 3-D imaging and stereographic projection are probably two entirely
different principles. I would appreciate any references (articles,
books, websites) that will help me to understand the concepts.

Thanks for any assistance you can provide!

Steve
--
Stephen J. Beck
Professor
Bio-Imaging Center/Electron Microscopy
Department of Biology
Nassau Community College
Garden City, NY 11530
Voice Mail: (516) 572-7829
Email: {becks-at-sunynassau.edu}
URL: {http://www.sunynassau.edu/webpages/biology/becks.htm}


From MicroscopyL-request-at-ns.microscopy.com Sun May 2 15:13:18 2004



From: yuansheng-sun-at-uiowa.edu (by way of Ask-A-Microscopist)
Date: Sun, 2 May 2004 15:16:54 -0500
Subject: [Microscopy] AskAMicroscopist: deconvolution for 3D phase contrast microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (yuansheng-sun-at-uiowa.edu) from http://www.msa.microscopy.com/Ask-A-Microscopist.html on Sunday, May 2, 2004 at 14:45:18
---------------------------------------------------------------------------

Email: yuansheng-sun-at-uiowa.edu
Name: Yuansheng Sun

Organization: University of Iowa

Education: Graduate College

Location: Iowa city, IA, USA

Question: Dear Madam or Sir:

I am doing a project on deconvolution for 3D phase contrast microscopy. As you know, it is very hard to measure a real psf. So I am wondering how to construct a theoretical psf based on the experimental parameters. Could you give me some infomation on this like some books or papers or technology documents. Thanks!
best from yuansheng

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sun May 2 16:07:25 2004



From: KEVIN W ELICEIRI :      eliceiri-at-facstaff.wisc.edu
Date: Sun, 02 May 2004 18:21:28 -0500
Subject: [Microscopy] Re: 3-D Imaging & Projection

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have been using ImageJ and the VolumeJ plugin (for surface
rendering) to generate 3D anaglyphs (red/green stereo) as animations.
As has been mentioned in this group before, ImageJ is a free program
that has been ported from NIH-Image to the Java platform so that it
will work on almost any machine. You can get more information on
ImageJ at http://rsb.info.nih.gov/ij/

You can download a couple of the anaglyphs at
http://astro.temple.edu/~jbs/research/Anaglyphs. Remember that you
will need red/green glasses to get the full effect.

Joel



The images were originally obtained as z-series with a Leica confocal
microscope.

Date sent: Sun, 02 May 2004 14:17:59 -0400
} From: Steve Beck {becks-at-sunynassau.edu}

I second Joel's suggestion of ImageJ/VolumeJ. You might also want to take a look at VisBio (http://www.loci.wisc.edu/VisBio). VisBio is a open-source JAVA based image analysis tool for multidimensional biological image data. Features include measurement, volume rendering, and arbitrary image stack slicing. Integration with ImageJ is under development.

best regards,
kevin

Kevin W. Eliceiri
Laboratory for Optical and Computational Instrumentation
http://www.loci.wisc.edu
Room 271 Animal Sciences
1675 Observatory Drive
Madison, WI 53706
Phone: 608-263-6288
Fax: 608-265-3083



From MicroscopyL-request-at-ns.microscopy.com Sun May 2 19:05:12 2004



From: Rosemary White :      Rosemary.White-at-csiro.au
Date: Mon, 03 May 2004 10:07:43 +1000
Subject: [Microscopy] Re: 3-D Imaging & Projection

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Steve,

I guess you know about the GeoWall (http://geowall.geo.lsa.umich.edu/)
projects. This is for geologists, but quite amenable to any type of 3D
projection really. Somewhat different to the setup you're proposing,
because it's PC- or Mac-based, but same principles. There's even a company
(VRCO) that sells turnkey systems. There is a bit of background information
on the GeoWall site, but more info on some of the link sites.

cheers,
Rosemmary



Dr. Rosemary White rosemary.white-at-csiro.au
Microscopy Centre ph. 61-2-6246 5475
CSIRO Plant Industry mob. 61-0402 835 973
GPO Box 1600 fax. 61-2-6246 5000
Canberra, ACT 2601
Australia


} From: Steve Beck {becks-at-sunynassau.edu}
} Date: Sun, 02 May 2004 14:17:59 -0400
} To: Microscopy-at-MSA.Microscopy.Com
} Subject: [Microscopy] 3-D Imaging & Projection
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Dear Colleagues,
}
} I am involved with a consortium that is being partially funded by the
} NSF-ATE program. The main project involves the development of 3-D
} images related to various science, math, engineering and technology
} areas (from molecules to organs to CAD/architecture) so that they can
} be projected (stereographic projection). Each institution
} participating in the project will have a 3-D theatre built so that
} SMET classes can be brought in for demonstrations of course specific
} images (the theatre will house a Barco 3-D projector driven by a SGI
} UNIX-based workstation and a special screen that retains
} polarization). Students and faculty from the partner institutions
} will also be trained in a residential program at Brookhaven National
} Lab by their 3-D Visualization Team.
}
} Since I have been teaching TEM and SEM courses at NCC for almost 20
} years, I thought that it would be an interesting project to render
} student and my own SEM & TEM images for 3-D projection. I have seen
} many threads on 3-D imaging on this listserver and I would appreciate
} some advice on where to begin. Initially, I am supposed to identify
} some software applications that would be suitable for rendering 3-D
} images of my micrographs. I am aware of some of the applications that
} are used at BNL for their 3-D theatre such as OpenDX, Open GL and
} Visualization ToolKit (VTK).
}
} Can anyone suggest an application (UNIX based ideally) that I might
} get started with? As a novice in this area I realize that the concept
} of 3-D imaging and stereographic projection are probably two entirely
} different principles. I would appreciate any references (articles,
} books, websites) that will help me to understand the concepts.
}
} Thanks for any assistance you can provide!
}
} Steve
} --
} Stephen J. Beck
} Professor
} Bio-Imaging Center/Electron Microscopy
} Department of Biology
} Nassau Community College
} Garden City, NY 11530
} Voice Mail: (516) 572-7829
} Email: {becks-at-sunynassau.edu}
} URL: {http://www.sunynassau.edu/webpages/biology/becks.htm}
}
}



From MicroscopyL-request-at-ns.microscopy.com Mon May 3 08:17:48 2004



From: Brian McIntyre :      mcintyre-at-optics.rochester.edu
Date: Mon, 3 May 2004 09:21:08 -0400 (EDT)
Subject: [Microscopy] Review of student projects

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all-

I've just completed a great semester of teaching EM techniques. The
student projects that were required are quite interesting.
If you have a chance, please review the student's work at this URL:

http://xray.optics.rochester.edu/workgroups/cml/opt307/index.html

Thanks!
Brian


_________________________________________________
Brian McIntyre
Univ. of Rochester
The Institute of Optics
River Campus EMLab
585-275-3058/4875
585-244-4936 (fax)




From MicroscopyL-request-at-ns.microscopy.com Mon May 3 08:36:45 2004



From: Lucian Barbu :      lbarbu-at-hasdeu.ubbcluj.ro (by way of
Date: Mon, 3 May 2004 08:40:21 -0500
Subject: [Microscopy] AskAMicroscopist: filament problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

Please help me. I'm really in trouble.
Just burned the second filament after 3 hours. Previously, I had about 200 h. Now the current is very unstable.
Before burning, is slowly going down and after that a quick increase up to 170 uA and bang.
We have a Jeol 1010 (1 year old).

Many thanks,
Luci
_________________________________________
Lucian Barbu-Tudoran
Electron Microscopy Center
Faculty of Biology & Geology
Babes-Bolyai University of Cluj-Napoca
5-7 Clinicilor Str., 3400 Cluj-Napoca, Romania
Tel/Fax: +40 264 598700
{http://hasdeu.ubbcluj.ro/emc/lucian_barbu_tudoran.htm} http://hasdeu.ubbcluj.ro/emc/lucian_barbu_tudoran.htm


From MicroscopyL-request-at-ns.microscopy.com Mon May 3 08:43:05 2004



From: mcintyre-at-optics.rochester.edu (by way of MicroscopyListserver)
Date: Mon, 3 May 2004 08:46:40 -0500
Subject: [Microscopy] viaWWW: Review of Student Projects

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (mcintyre-at-optics.rochester.edu) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Monday, May 3, 2004 at 08:10:50
---------------------------------------------------------------------------

Email: mcintyre-at-optics.rochester.edu
Name: Brian McIntyre

Organization: Univ.of Rochester

Title-Subject: [Microscopy] [Filtered] Review of Student Projects

Question: Hi all-

I've just concluded a great semester of teaching EM techniques to a mixed group of undergrads and grads. They've put together some interesting projects that are posted on the web here:

http://xray.optics.rochester.edu/workgroups/cml/opt307/index.html

Please take a look and provide feedback to them and me.

Thanks!
Brian

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon May 3 09:12:16 2004



From: Zhang, Guofeng (NIH/OD/ORS) :      zhangguo-at-ors.od.nih.gov
Date: Mon, 3 May 2004 10:15:19 -0400
Subject: [Microscopy] TEM: Membrane Visualization Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


} Dear Listserver,
}
} I have recently processed Hela cells grown on covertslips by
} conventional chemical fixation( 2.5% formaldehyde plus 2.5% glutaraldehyde
} for 2h and postfixation in 1.5% Osmium in 0.1M sodium cacodylate buffer,
} dehydrated in a series of concentrations of ethanol over a period of 90
} minutes and then infiltrated in Epon-aradite). After sectioning, I stained
} the sections with 2% uranyl acetate in distill water for 10 min and lead
} citrate for 5 min. I could hardly visualize the membrane structrue,
} particularly nuclear membrane, plasma membrane, Golgi apparatus and ER
} membranes. However the other structures were well preserved. I am here
} asking for your advice and greatly appreciate any suggestions!
}
} Thank you very much in advance!
}
} Guofeng Zhang, Ph.D
} Division of Bioengineering & Physical Sciences
} National Institute of Health
} Bethesda, MD 20892
} Tel: 301-451-3856
} Email: zhangguo-at-mail.nih.gov
}
}


From MicroscopyL-request-at-ns.microscopy.com Mon May 3 12:14:15 2004



From: Diane.Ciaburri-at-gd-ais.com
Date: Mon, 3 May 2004 13:16:22 -0400
Subject: [Microscopy] EDS of phosphors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello,

I'm trying accquire EDS spectra of various posphors and am having
difficulty. There are peaks that do not seem to match any known elements,
and all of the peaks shift up and down in energy from one spectrum to the
next. I'm wondering if this is a common phenomenon with phosphors, or is my
system out to lunch? All ideas are welcome!

Diane Ciaburri


From MicroscopyL-request-at-ns.microscopy.com Mon May 3 12:33:56 2004



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Mon, 03 May 2004 12:36:20 -0500
Subject: [Microscopy] Re: TEM: Membrane Visualization Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Guofeng

Having been around for what seems like forever, I really did not write
up some of the old literature as some of our students have suggested.
But there are old reports of just this problem. The conclusion was that
osmication does not adequately stabilize membranes against extraction
during ethanol dehydration. The reference is a bit fuzzy, but I believe
it was Dan Pease’s book ‘Histological Techniques for Electron
Microscopy’, 2nd edition, 1964. It was also never clear whether it was
the ethanol or the propylene oxide intermediate which was responsible
for membrane extraction. We have tested this out and confirmed that,
for whatever reason, in our hands we will see the loss of membranes you
describe if we use the procedures as you list.

There are two solutions. First, post-fixation in uranyl acetate.
Uranyl acetate will stabilize membranes. We wash 1x in buffer and 3x in
filtered water after the osmication step, and then fix in 2% UA. Our
monolayers and free suspensions are usually only fixed for a couple of
hours, but we do leave them in UA for up to several days without noting
any deleterious effect. This step usually improves stabilization of
membranes in our hands.

The second solution is to dehydrate in acetone. There are two reasons,
first is that acetone is not reported to extract membranes, and second,
you do not need an intermediate step such as propylene oxide, but go
directly into graded acetone:plastic for infiltration. This second
reason saves on costs, reduces turnaround, and is a little more
environmentally friendly. In fact, we have found that using acetone is,
by itself, sufficient to ensure good membrane stabilization.

Our normal procedure now is to use both UA en bloc postfixation followed
by acetone dehydration. Our results are usually very good, and we never
lose membranes.

Hope these ideas help.

Paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-954
Cell:204-781-1502
Fax:204-789-3926



From MicroscopyL-request-at-ns.microscopy.com Mon May 3 13:07:05 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 3 May 2004 11:14:57 -0700
Subject: [Microscopy] Re: 3-D Imaging & Projection

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 2, 2004, at 11:17 AM, Steve Beck wrote:

} Can anyone suggest an application (UNIX based ideally) that I might
} get started with? As a novice in this area I realize that the concept
} of 3-D imaging and stereographic projection are probably two entirely
} different principles. I would appreciate any references (articles,
} books, websites) that will help me to understand the concepts.
}
Dear Steve,
If you do not get sufficient info from this list, try the 3dem list:

} ADMINISTRIVIA
}
} To post to this list, send mail to
}
} 3dem-at-ucsd.edu
}
} To subscribe to this list, send mail to
}
} majordomo-at-3dem.ucsd.edu
}
} with the text
}
} subscribe 3dem
}
} in the body of the message (not in the Subject:).
}
} To unsubscribe from this list, send mail to
}
} majordomo-at-3dem.ucsd.edu
}
} with the text
}
} unsubscribe 3dem
}
} in the body of the message (not in the Subject:).

I also know that the group at the Wadsworth Center in Albany NY has
prepared stereo images from 3-D reconstructions, so you might want to
contact them.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon May 3 13:38:17 2004



From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Mon, 3 May 2004 08:40:46 -1000 (HST)
Subject: [Microscopy] Re: TEM: Membrane Visualization Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} } I have recently processed Hela cells grown on covertslips by
} } conventional chemical fixation( 2.5% formaldehyde plus 2.5% glutaraldehyde
} } for 2h and postfixation in 1.5% Osmium in 0.1M sodium cacodylate buffer,
} } dehydrated in a series of concentrations of ethanol over a period of 90
} } minutes and then infiltrated in Epon-aradite). After sectioning, I stained
} } the sections with 2% uranyl acetate in distill water for 10 min and lead
} } citrate for 5 min. I could hardly visualize the membrane structrue,
} } particularly nuclear membrane, plasma membrane, Golgi apparatus and ER
} } membranes. However the other structures were well preserved. I am here
} } asking for your advice and greatly appreciate any suggestions!

Try increasing the time in uranyl acetate to 25 or more minutes, and
decrease the time in lead citrate to 3 minutes. Unintuitively, sometimes
longer times in the lead citrate causes bleaching rather than staining!

On the other hand, I've also had success with 10 minutes in each, and with
5 minutes in lead followed by 10 minutes in UA followed by another 10
minutes in lead, and many other variations. Try several approaches and see
what works for you!

Aloha,
Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************



From MicroscopyL-request-at-ns.microscopy.com Mon May 3 14:08:00 2004



From: JoAnn Buchanan :      redhair-at-stanford.edu
Date: Mon, 03 May 2004 12:11:11 -0700
Subject: [Microscopy] TEM: Membrane Visualization Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Guofeng. I have worked on many cell culture preps. I normally use
microwave processing however for bench processing I would recommend that
you shorten all your processing steps. I would bench fix in glut and
osmium for 20 mins each at room temperature for example. I use 2%
glutaraldehyde on 0.1 M cacodylate buffer. I wash in 0.1M cacodylate
buffer with 0.3M sucrose added. I also add 0.3 M sucrose to the post fix.
The cells are very thin and not like a tissue chunk. Also, dehydration
times should be very short- 1-2 mins is sufficient. in your alcohol changes.
Cultured cells don't have much contrast so try using a reduced osmium fix
(2% osmium tetroxide in 0.1 M cacodylate buffer containing 0.8% potassium
ferricyanide). it gives really nice contrast to membranes and
cytoskeleton. Also try an en bloc UA step before the dehydration.
Your post staining sounds OK.
Good luck
JoAnn Buchanan
Dear Listserver,
}
} I have recently processed Hela cells grown on covertslips by
} conventional chemical fixation( 2.5% formaldehyde plus 2.5% glutaraldehyde
} for 2h and postfixation in 1.5% Osmium in 0.1M sodium cacodylate buffer,
} dehydrated in a series of concentrations of ethanol over a period of 90
} minutes and then infiltrated in Epon-aradite). After sectioning, I stained
} the sections with 2% uranyl acetate in distill water for 10 min and lead
} citrate for 5 min. I could hardly visualize the membrane structrue,
} particularly nuclear membrane, plasma membrane, Golgi apparatus and ER
} membranes. However the other structures were well preserved. I am here
} asking for your advice and greatly appreciate any suggestions!
}
} Thank you very much in advance!
}
} Guofeng Zhang, Ph.D
} Division of Bioengineering & Physical Sciences
} National Institute of Health
} Bethesda, MD 20892
} Tel: 301-451-3856
} Email: zhangguo-at-mail.nih.gov
}


Department of Molecular and Cellular Physiology
Stanford University School of Medicine
Stanford, CA 94305
650-723-5856



From MicroscopyL-request-at-ns.microscopy.com Mon May 3 14:24:13 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 3 May 2004 12:32:04 -0700
Subject: [Microscopy] Re: EDS of phosphors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 3, 2004, at 10:16 AM, Diane.Ciaburri-at-gd-ais.com wrote:

} I'm trying accquire EDS spectra of various posphors and am having
} difficulty. There are peaks that do not seem to match any known
} elements,
} and all of the peaks shift up and down in energy from one spectrum to
} the
} next. I'm wondering if this is a common phenomenon with phosphors, or
} is my
} system out to lunch? All ideas are welcome!
}
Dear Diane,
I assume you're looking at inorganic phosphors, such as CdS, etc., in
which case there should be no problem. Just because it's a phosphor
does not change the interaction with beam electrons that produces
x-rays. There might be contaminants in the mix, but these would be
minor--not ~10%--so there could be some very small peaks, but
everything should match one or more known elements. Have you
calibrated the energy spectrum? The energies should not shift--the
only way that might happen is if the majority of the x-rays were
inelastically scattered. If you take multiple spectra from a single
area (assuming that there is not significant loss of material during
data collection) neither the energies nor the intensities should vary.
I suggest putting in a standard, such as Al evaporated onto a formvar
film on a Cu grid, and taking spectra from locations where both the Al
and Cu signals are strong. The spectra should be consistent, and you
should be able to check the energy calibration and see if the
resolution is OK. To see if the resolution meets spec, put in a grid
with Mn evaporated on formvar and measure the FWHM of the peak.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon May 3 15:03:34 2004



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Mon, 3 May 2004 16:59:34 -0400
Subject: [Microscopy] Re: TEM: Membrane Visualization Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Try changing your fixation protocols. I use the following fixatives
to help with membrane preservation:

"Yellow Fix for EM" (based on Ito, S and Karnovsky J. J Cell. Biol
Abstract 1968)
2.5% glut.
4.0% paraformaldehyde
0.02 % picric acid
in 0.1M sodium cacodylate buffer

(I use 20ml of 0.2M buffer, 10ml of 10% glut, 10 ml of 16% pfa and 2
ml of a saturated sol. of picric acid--I keep a jar of it in my hood,
with the crystals covered with water and draw off as I need it,
adding fresh water when I'm done so that it doesn't explode.)

For a secondary fix I use 1% osmium tetroxide and 1.5% potassium
ferricyanide (aqueous)made up my mixing stock solutions of 2% osmium
and 3% K-ferr 1:1.

Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Mon May 3 15:06:27 2004



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Mon, 03 May 2004 15:09:37 -0500
Subject: [Microscopy] Re: EDS of phosphors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Cool! Well, maybe not for you.

We have an infrared chamber-scope that I have to instruct users to turn off
before they do any EDS. Otherwise, the flood of IR photons overwhelms the
EDS detector and runs it up to 100% dead time. Under less overwhelming
conditions, the photons are still detected and their energy adds on to the
real energy of the x-ray photons and peaks get shifted upscale. Its kind of
fun to watch the effect on our spectrum as we turn the light on and off and
the spectral peaks shift back and forth.

I had never thought about it, but the same effect should be expected when
working with cathodo-luminescent materials. And your phosphors are
definitely luminescent.

I recall that some EDS detectors have their windows coated (with Al, I
think) to avoid just this problem. I don't know if that is something that
can be done very well to an existing detector. I don't suppose you could
interpose another film in front of your detector; it would probably also
absorb your x-rays.

I wonder how others would suggest dealing with this.

Warren

At 12:16 PM 5/3/2004, you wrote:
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

-------------------------------------------
No files should be attached to this message
-------------------------------------------
Warren E. Straszheim, Ph.D.
Materials Analysis and Research Lab
Iowa State University
46 Town Engineering
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
Web: www.marl.iastate.edu

Scanning electron microscopy, x-ray analysis, and image analysis of materials
Computer applications and networking



From MicroscopyL-request-at-ns.microscopy.com Mon May 3 15:30:18 2004



From: Kestutis Smalinskas :      smalinskas-at-yahoo.com
Date: Mon, 3 May 2004 13:33:30 -0700 (PDT)
Subject: [Microscopy] AskAMicroscopist: filament problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Luci, the Microscopy Today journal, March/April 2003
issue has an article on "The Life and Death of a
Tungsten Hairpin Filament". In it there is described
the different failure modes for a hairpin filament
which may help you.

You should be able to access this article via the
following link:

http://www.microscopy-today.com/TableofContentsPDF.html

Stu Smalinskas, P.E.
Metallurgist
SKF
46815 Port Street
Plymouth, Michigan 48170

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

Luci wrote:

Hi

Please help me. I'm really in trouble.
Just burned the second filament after 3 hours.
Previously, I had about 200 h. Now the current is very
unstable.
Before burning, is slowly going down and after that a
quick increase up to 170 uA and bang.
We have a Jeol 1010 (1 year old).

Many thanks,
Luci
_________________________________________
Lucian Barbu-Tudoran
Electron Microscopy Center
Faculty of Biology & Geology
Babes-Bolyai University of Cluj-Napoca
5-7 Clinicilor Str., 3400 Cluj-Napoca, Romania
Tel/Fax: +40 264 598700
{http://hasdeu.ubbcluj.ro/emc/lucian_barbu_tudoran.htm




__________________________________
Do you Yahoo!?
Win a $20,000 Career Makeover at Yahoo! HotJobs
http://hotjobs.sweepstakes.yahoo.com/careermakeover


From MicroscopyL-request-at-ns.microscopy.com Mon May 3 16:24:54 2004



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Mon, 03 May 2004 15:27:00 -0700
Subject: [Microscopy] Re: Re: TEM: Membrane Visualization Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We would guess the problem may be in your power supply. Your 200+ hours
indicates a good vacuum and clean system. Our customers average 200+ on
rebuilt filaments and slightly less on new filaments due to the special
process used on rebuilt filaments.

Based on your comments, I'd check with Jeol services.

John Arnott

Disclaimer: Ladd Research is a supplier of a wide variety of microscope
supplies and accessories, including new and rebuilt filaments.


Ladd Research
83 Holly Court
Williston, VT 05495

On-line Catalog: http://www.laddresearch.com

tel: 1-802-658-4961(anywhere) or 1-800-451-3406(US)
fax: 1-802-660-8859
e-mail: sales-at-laddresearch.com


----- Original Message -----
} From: "Lucian Barbu (by way of Ask-A-Microscopist)"
{lbarbu-at-hasdeu.ubbcluj.ro}
To: {microscopy-at-ns.microscopy.com}
Sent: Monday, May 03, 2004 9:40 AM

Bill,

Yes, the system was just calibrated last Friday, and I double-checked it
again today which led me to think there must be something funny happening
with the samples. Here's the answer, thanks to a variety of helpful hints I
received. Apparently my detector is sensitive to visible light and the
whole spectrum was being shifted to higher energies by varying amounts
(depending on how much light I was generating). By switching from my thin
window to my berillium window, and blocking the visible light, things make
sense again.

Thanks, Bill and everyone else that responded. You guys are life savers!

Diane Ciaburri


-----Original Message-----
} From: Bill Tivol [mailto:tivol-at-caltech.edu]
Sent: Monday, May 03, 2004 3:32 PM
To: microscopy-at-msa.microscopy.com

Dear Diane,
If you are using the modern thin-window EDS detectors sold nowadays, you should
know that they are very susceptible to light and that could easily be the source
of your strange signals. The old-style Be-window detectors were not
light-susceptible, so if you can find one you can test the elements above Ne.
For the elements below that, perhaps using a low accelerating voltage ( {5 kV)
will reduce the light emission from the phosphor enough to get a reading on a
thin-window detector.
Regards,
Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca
----- Original Message -----
} From: {Diane.Ciaburri-at-gd-ais.com}
To: {Microscopy-at-MSA.Microscopy.Com}
Sent: Monday, May 03, 2004 10:16 AM

Paul and Guofeng
I never though for staining in UA after osmication as a
"postfixation". It's hardly believe to me that UA has something to do with
"fixation". In my point of view UA stained positively nucleic acids and
don't interfere with membranes. When I need reduce the appearance of the
ribosomes in neurone's cytoplasm, I used to remove UA staining after
osmication step from procedure. In my hands, it absolutely does not affect
membranes (I did both: with and without that step). I used Spurr, not
Epon. I also did not see any effect of ethanol on membranes. I think, if
you are using it for 10-15 min each step, it should not affect membrane's
quality, otherwise nearly whole EM is pure artefact (most people still use
ethanol). The basic reason, not to use acetone for dehydratation is
because acetone is very strong denaturing agent (contrary to ethanol), so
acetone will definitely alternate the fine structure. For this reason,
people use acetone in freeze-substitution procedure at low temperature to
reduce denaturing effect of acetone. Returning back to original question,
I would suggest to check is OsO4 was good or not. If there is any
suspicious there, you need to use fresh solution/chemicals. After
osmication the tissue (even in monolayer) should turn brown. Sergey

At 12:36 PM 5/3/2004 -0500, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon May 3 18:28:33 2004



From: donaldawbrey-at-hotmail.com (by way of MicroscopyListserver)
Date: Mon, 3 May 2004 21:50:24 -0500
Subject: [Microscopy] viaWWW: EM scopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Mary & Co.,

I think Diane's issue with phosphorous is visible fluorescence similar to the face of a CRT. After all, to get an image on a TV screen we shoot electrons at phosphorous coated screens and hence it fluoresces in visible light, and likely IR as well. Ideal to make the EDX detector do odd things. Chamberscopes, as Warren indicated, have the same effect.

Peter Tomic
Agere Systems

-----Original Message-----
} From: Mary Mager [mailto:mager-at-interchange.ubc.ca]
Sent: Monday, May 03, 2004 5:53 PM
To: Diane.Ciaburri-at-gd-ais.com
Cc: Microscopy

Dear Diane,
If you are using the modern thin-window EDS detectors sold nowadays, you should
know that they are very susceptible to light and that could easily be the source
of your strange signals. The old-style Be-window detectors were not
light-susceptible, so if you can find one you can test the elements above Ne.
For the elements below that, perhaps using a low accelerating voltage ( {5 kV)
will reduce the light emission from the phosphor enough to get a reading on a
thin-window detector.
Regards,
Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca
----- Original Message -----
} From: {Diane.Ciaburri-at-gd-ais.com}
To: {Microscopy-at-MSA.Microscopy.Com}
Sent: Monday, May 03, 2004 10:16 AM

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (donaldawbrey-at-hotmail.com) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Monday, May 3, 2004 at 12:44:21
---------------------------------------------------------------------------

Email: donaldawbrey-at-hotmail.com
Name: Donald G. Awbrey HT(ASCP) QIHC

Organization: Harris Methodist Hospital

Title-Subject: [Microscopy] [Filtered] MListserver: EM scopes

Question: Dear MSA netters,

Our department is in the preliminary stages of aquiring a new electron microscope. The one in use is antiquated. I would like to have as many opinions of different scopes and their ancillary systems.

We are looking for a scope to be used for diagnostic services (such as kidney biopsies, tumors, and etc.).

Our requirements are: 120kv, flat bed camera scanner, subliminal printer, image software, dual digital/manual capabilities.

Thank you in advance for any opinions.


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon May 3 21:47:58 2004



From: ecd10-at-psu.edu (by way of MicroscopyListserver)
Date: Mon, 3 May 2004 21:51:18 -0500
Subject: [Microscopy] viaWWW: Postdoc opening at Penn State

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (ecd10-at-psu.edu) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Monday, May 3, 2004 at 14:10:00
---------------------------------------------------------------------------

Email: ecd10-at-psu.edu
Name: Elizabeth Dickey

Organization: Pennsylvania State University

Title-Subject: [Microscopy] [Filtered] MListserver:Postdoc opening at Penn State

Question: POST-DOCTORAL POSITION
in
Electron Microscopy of Nanostructured Electronic Materials
at
The Pennsylvania State University



A postdoctoral position is available in the area of nanostructured electronic materials beginning July 1, 2004. The research project focuses on understanding interface structure and chemistry in heterostructured thin films and nanowires. Aspects of the project will be conducted under the auspices of an NSF-NIRT program on semiconducting nanowires. Through a variety of electron imaging and spectroscopy techniques we aim to quantify atomic structure and chemistry of interfaces in heterostructured materials. Most of the research will be conducted on a JEOL 2010F Field Emission TEM/STEM outfitted with an EDAX energy dispersive x-ray spectrometer (EDS), Gatan Enfina electron energy loss spectrometer (EELS), high-angle annular dark field STEM detector, and acquisition hardware and software for spectrum imaging. The ideal candidate for this position will have experience in HRTEM, STEM, EDS and EELS. The salary will be commensurate with qualifications and experience.

Please forward questions or send applications to:

Professor Elizabeth Dickey
Department of Materials Science and Engineering
The Pennsylvania State University
223 Materials Research Building
University Park, PA 16802
USA

tel: (814) 865-9067
FAX: (814) 865-2326
email: ecd10-at-psu.edu


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue May 4 03:20:46 2004



From: sylvain maury :      sylvain.maury-at-thalesgroup.com
Date: Tue, 04 May 2004 10:23:57 +0200
Subject: [Microscopy] syton european distributors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all !


does anyone know some manufacturers or distributors of coloïdal silica
also called "syton", for microsectionned dies polishing process ?

thanx in advance...

Sylvain MAURY



From MicroscopyL-request-at-ns.microscopy.com Tue May 4 11:03:10 2004



From: David Henriks :      henriks-at-southbaytech.com
Date: Tue, 04 May 2004 09:14:07 -0700
Subject: [Microscopy] Re: syton european distributors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Sylvain:

South Bay Technology offers 3 types of Colloidal Silica for cross
sectioning of semiconductor materials - as well as other applications.
The products we offer are:

CS1 - Non-crystallizing Colloidal Silica
CS2 - Syton
CS3 - Glanzox

All of these products are available through our local distributor in
France. All of our distrinbutor contact information can be found on our
website. As luck would have it, "Colloidal Silica" is also the featured
product on our website today. You can go to www.southbaytech.com and
find information on the colloidal silica products on the main page.
Alternatively, you can enter "colloidal silica" in the product search
box to bring up all references to our various colloidal silica products.

I hope this helps.

Best regards-

David

sylvain maury wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
David Henriks
Vice President

South Bay Technology, Inc.
1120 Via Callejon
San Clemente, CA 92673 USA

TEL: +1-949-492-2600
Toll-free in the USA: +1-800-728-2233
FAX: +1-949-492-1499

email: henriks-at-southbaytech.com

Celebrating 40 years of providing Materials Processing Solutions for Metallogaphy, Crystallography and Electron Microscopy.

Please visit us online at www.southbaytech.com.

The information contained in this message and any attachments is privileged and confidential. This message is intended for the individual or entity addressed.
If you are not the intended recipient, please do not read, copy or disclose this communication. Notify the sender of the mistake by calling +1-949-492-2600 and
delete this message from your system.





From MicroscopyL-request-at-ns.microscopy.com Tue May 4 12:41:00 2004



From: sara goldston :      cogold13-at-yahoo.com
Date: Tue, 4 May 2004 10:43:46 -0700 (PDT)
Subject: [Microscopy] Re: viaWWW: EM scopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The clinical lab I used to work in had a Codonex
di-sub printer which we rarely used. Overall, our
imaging software (Gatan Micrograph ~2+years old) did
not jive with the printer too well - less resolution
with printed image. Not to say its a bad system. It
might have been because our imaging program needed
updating. Case in point, scanned images that were
'tweeked' in Adobe and printed with the Codonex came
out excellent.
Sara Goldston

by way of MicroscopyListserver
{donaldawbrey-at-hotmail.com} wrote:

------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy
Society of America

Below is the result of your feedback form
(NJZFM-ultra-55). It was submitted by
(donaldawbrey-at-hotmail.com) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html
on Monday, May 3, 2004 at 12:44:21
---------------------------------------------------------------------------

Email: donaldawbrey-at-hotmail.com
Name: Donald G. Awbrey HT(ASCP) QIHC

Organization: Harris Methodist Hospital

Title-Subject: [Microscopy] [Filtered] MListserver: EM
scopes

Question: Dear MSA netters,

Our department is in the preliminary stages of
aquiring a new electron microscope. The one in use is
antiquated. I would like to have as many opinions of
different scopes and their ancillary systems.

We are looking for a scope to be used for diagnostic
services (such as kidney biopsies, tumors, and etc.).

Our requirements are: 120kv, flat bed camera scanner,
subliminal printer, image software, dual
digital/manual capabilities.

Thank you in advance for any opinions.


---------------------------------------------------------------------------






__________________________________
Do you Yahoo!?
Win a $20,000 Career Makeover at Yahoo! HotJobs
http://hotjobs.sweepstakes.yahoo.com/careermakeover


From MicroscopyL-request-at-ns.microscopy.com Tue May 4 16:31:39 2004



From: Zhang, Guofeng (NIH/OD/ORS) :      zhangguo-at-ors.od.nih.gov
Date: Tue, 4 May 2004 17:34:35 -0400
Subject: [Microscopy] Many thanks to all responses to Membrane Visualization Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Listserver,

Many thanks to all of your helpful comments and suggestions on membrane
visualization problems! I think I will get succeed in solving the problems
in near future after taking your advise.

Kind Regards!

Guofeng
}
} Guofeng Zhang, Ph.D
} Division of Bioengineering & Physical Sciences
} National Institute of Health
} Bethesda, MD 20892
} Tel: 301-451-3856
} Email: zhangguo-at-mail.nih.gov
}
}


From MicroscopyL-request-at-ns.microscopy.com Tue May 4 17:14:38 2004



From: Bill Chissoe :      wchiss-at-ou.edu
Date: Tue, 04 May 2004 17:18:52 -0500
Subject: [Microscopy] Seed prep

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Listers,
I have a researcher wanting to look at seed surface morphology. These
seeds are from dried herbarium sheets. I initially had him simply treat
the seeds with acetone to remove any waxes from the surface and then
O-T-O-T-O them. This was not adequate as we saw what I believe to be
fungal hyphae and spores plus a LOT of debris. Does anybody have a
"tried and true" (and easy!) method of preparing seeds? I was thinking
about doing acetolysis or 10% KOH, which I do with pollen routinely, but

wasn't sure how the seed surface would hold up. Also, I don't think the
person sonicated in acetone. Might that alone be adequate? TIA.

Bill

--
=============================================================
Bill Chissoe III
Electron Microscopist
University of Oklahoma
770 Van Vleet Oval
Norman, Ok. 73019
E-mail: wchiss-at-ou.edu Ph. (405)325-4391
=============================================================




From MicroscopyL-request-at-ns.microscopy.com Tue May 4 20:01:43 2004



From: cmeyer911-at-yahoo.com (by way of MicroscopyListserver)
Date: Tue, 4 May 2004 20:05:09 -0500
Subject: [Microscopy] viaWWW: [TEM EELS Gatan 607

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (cmeyer911-at-yahoo.com) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, May 4, 2004 at 19:31:16
---------------------------------------------------------------------------

Email: cmeyer911-at-yahoo.com
Name: Chris Meyer

Organization: Boeing

Title-Subject: [Microscopy] [Filtered] [TEM EELS Gatan 607]

Question: I have an old Gatan Model 607 EELS system that I have never run (I'm new to TEM). It used to be run by a TN-5500 system, that is soon to be replaced with a new EDS system. I do not want to lose EELS capability for future projects. A Gatan representative informed me the 607 system could be run with a Mac with an old operating system.

I'd like to talk (email) to any current or former 607 EELS users that would have ideas on how I could keep this capability for the future. Is my only option to keep the TN-5500 for this purpose alone? I'd appreciate it. Thanks.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed May 5 02:58:16 2004



From: =?iso-8859-1?Q?KJ_H=FCbner?= :      hubner-at-IOd.krakow.pl
Date: Wed, 5 May 2004 09:36:51 +0200
Subject: [Microscopy] Re: syton european distributors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

look on the www.struers.com

best regards

chris


From MicroscopyL-request-at-ns.microscopy.com Wed May 5 07:49:16 2004



From: jacqui.ross-at-auckland.ac.nz (by way of MicroscopyListserver)
Date: Wed, 5 May 2004 07:52:32 -0500
Subject: [Microscopy] viaWWW: inverted fluorescence microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (jacqui.ross-at-auckland.ac.nz) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, May 4, 2004 at 23:18:40
---------------------------------------------------------------------------

Email: jacqui.ross-at-auckland.ac.nz
Name: Jacqui Ross

Organization: The University of Auckland

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: We are about to buy a new inverted fluorescence microscope, probably a Nikon.

Does anyone have any experience/feedback about the Nikon Digital Sight colour camera (DS-5Mc-U1)? We are intending to buy the new cooled one and I would appreciate hearing from anyone who has used one or seen one demonstrated recently.

Cheers,

Jacqui.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed May 5 14:48:22 2004



From: Ron Anderson :      randerson20-at-tampabay.rr.com
Date: Wed, 5 May 2004 15:51:27 -0400
Subject: [Microscopy] Microscopy Today May Issue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Listers,

Here is the TABLE OF CONTENTS for the MAY 2004 issue of MICROSCOPY TODAY.


New Subscriptions via http://www.microscopy-today.com only, please.
New subscriptions will close on Tuesday May 11th for this issue.

WE HAVE PURGED NON-MSA MEMBER, NON-NORTH AMERICAN, UNSUBSCRIBED INDIVIDUALS!
SEVERAL HUNDRED NON-QUALIFIED SUBSCRIBERS HAVE BEEN PURGED.

Scanning Wet Specimens
.Stephen W. Carmichael and Wilma L. Lingle, Mayo Clinic

Laboratory Design for High-Performance Electron Microscopy
.M.A. O'Keefe, et al., LBNL; C.J.D. Heatherington, et al., Oxford; B.
.Carragher, et al., Scripps; L.F. Allard, et al., ORNL

New CMEIAS Image Analysis Software for Computer-Assisted Microscopy of
Microorganisms and Their Ecology
.Frank B. Dazzo, Michigan State University

Making Slide Shows in Acrobat
.Jerry Sedgewick, University of Minnesota

Sub-Angstrom Resolution with a Mid-Voltage TEM
.M.A. O'Keefe, C.J.D. Hetherington, E. C. Nelson, LBNL, Berkeley

Viability and Versatility of the Yeast Cell
.Michelle J. Henry-Stanley & Carol L. Wells, Univ. of Minnesota

New Adhesion Mechanism in Giardia: Role of the Ventrolateral Flange in the
Attachment of Trophozoites to Rough and Porous Surfaces
.S.L. Erlandsen, U. Minnesota; A.P. Russo, & J.N. Turner, New York
.Wadsworth Center

Project VISUAL: Facilitating the Connection Between Art and Science
.L.M. Strzegowski & T.P. Russell, U. of Massachusetts, Amherst

Use and Disposal of Uranyl Acetate in the Electron Microscope Laboratory:
Glow in the Dark or Walk in the Park?
.Randy Tindall, University of Missouri

Confocal Microscopy for Diagnostic Cytology
.M.E. Boon, & L.P. Kok, U. Groningen, The Netherlands

Technical Note on the Preparation of Un-decalcified Trabecular Bone for
Examination by TEM
.Jeannette Taylor, Emory Univ. & Iwona Jasiuk, Georgia Inst. of Technology

The Low Voltage SEM Imaging Advantage: A Reminder
.Steven S. Hurban. Endicott Interconnect Technologies, Inc. NY

Ron Anderson, Editor
Microscopy Today




From MicroscopyL-request-at-ns.microscopy.com Wed May 5 16:53:09 2004



From: fremingt-at-fhcrc.org (by way of MicroscopyListserver)
Date: Wed, 5 May 2004 16:56:40 -0500
Subject: [Microscopy] viaWWW: Many Thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (fremingt-at-fhcrc.org) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, May 5, 2004 at 11:24:55
---------------------------------------------------------------------------

Email: fremingt-at-fhcrc.org
Name: Franque Remington

Organization: Fred Hutchinson Cancer Research Center

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Many Thanks to all who offered suggestions and techniques for preserving conidia for SEM. They have been very helpful.

Franque Remington

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed May 5 16:53:47 2004



From: bwareham-at-utah.gov (by way of MicroscopyListserver)
Date: Wed, 5 May 2004 16:57:18 -0500
Subject: [Microscopy] viaWWW: Fixation of animal lung tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (bwareham-at-utah.gov) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, May 5, 2004 at 15:08:22
---------------------------------------------------------------------------

Email: bwareham-at-utah.gov
Name: Beverly Wareham

Organization: Utah Veterinary Diagnostic Lab

Title-Subject: [Microscopy] [Filtered] Fixation of animal lung tissue

Question: Hi all,
Does anyone have a protocol for fixing animal lung tissue (sheep) for EM and is willing to share it? I'd really appreciate it.

Thanks,
Beverly Wareham
Utah Veterinary Diagnostic Laboratory
435-797-1799
bwareham-at-utah.gov

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed May 5 16:54:19 2004



From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Wed, 5 May 2004 16:57:29 -0500
Subject: [Microscopy] Lab cleanup news (good)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear listers,

Sometimes you just gotta share the good news. After trying out my share
of handheld vacuum cleaners for small lab cleanup jobs and finding them
all to be pretty asthmatic, I finally hit paydirt. I found something
called the Euro Pro 700 watt hand vac down at the local Office Depot for
about $30, complete with small hose, attachments and shoulder strap.
Not only does it clean up the plastic fluff that builds up in
ultramicrotomy areas, it would probably suck up the whole microtome if
you turned your back. That's probably why it's also called "The Shark".

No financial interest, just a (finally) happy owner of a tiny vacuum
cleaner that works!

Randy Tindall
EM Specialist
Electron Microscopy Core Facility---We Do Small Well!
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/




From MicroscopyL-request-at-ns.microscopy.com Wed May 5 16:54:16 2004



From: ananthpb-at-rediffmail.com (by way of MicroscopyListserver)
Date: Wed, 5 May 2004 16:57:44 -0500
Subject: [Microscopy] viaWWW: electropolisher

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (ananthpb-at-rediffmail.com) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, May 5, 2004 at 16:07:20
---------------------------------------------------------------------------

Email: ananthpb-at-rediffmail.com
Name: Ananth Puthucode

Organization: Univ. North Texas

Title-Subject: [Microscopy] [Filtered] electropolisher

Question: Hi all,

I am looking for a used electropolisher for TEM sample preparation. Please let me know if anyone has got one and planning to sell it.

Thanks,
Ananth

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu May 6 03:53:57 2004



From: sylvain maury :      sylvain.maury-at-thalesgroup.com
Date: Thu, 06 May 2004 10:57:05 +0200
Subject: [Microscopy] method for polishing semiconductors with syton or colloidal silica

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

hello microscopists !

I have a question for those who work in Failure Analysis of
semiconductors but anyone can answer...

I'm trying to define a methodology to use colloidal silica for polishing
process of microsectionned dies (Si, AsGA), for SEM observations. I'd
like to know the conditions that people commonly use for their sample
preparations with Syton.
Any suggestions are welcome...
Thanx in advance!

and thanx to David Henricks : i just received the documents about
colloidal silica and SEM sample preparation. But the paper about iridium
sputtering was not in the package. You must have forgotten it, but it
doesn't matter. thanx again...

Sylvain MAURY



From MicroscopyL-request-at-ns.microscopy.com Thu May 6 07:28:23 2004



From: =?iso-8859-1?Q?KJ_H=FCbner?= :      hubner-at-IOd.krakow.pl
Date: Thu, 6 May 2004 14:08:17 +0200
Subject: [Microscopy] Fw: method for polishing semiconductors with syton or colloidal silica

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


-----


} hi !
}
} ask the company name "logitech" (a division from struers denmark)
} they are have a special equipment for polishing semiconductors !
} email info-at-logitech-us.com or www.logitech.uk.com
}
} best regards
}
} chris huebner
}
} Foundry Research Insitute, Kraków - Poland
}



From MicroscopyL-request-at-ns.microscopy.com Thu May 6 08:38:24 2004



From: Jeff Stewart :      jeff-at-metallography.com
Date: Thu, 6 May 2004 09:44:36 -0400
Subject: [Microscopy] IMS seeks Local Chair for 2005 event

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Randy

I don't know if anyone else would wish to confirm my fears but I am concerned about the use of any basic vacuum cleaner for clearing up resin dust.

The problem as I see it is:
1. Vacuum cleaners do pump out small particles from their exhaust (unless they have some very fine submicron exhaust filter).
2. It's the fine stuff (below 2 or 3 microns) that remains airborne and worse still is the ideal size for inhalation. It is much more likely to penetrate deeper into the respiratory tract (hence the problem with certain sizes of dust such as asbestos).
3. Resins used for most biological purposes are not completely polymerised at 60 to 80 degrees. They are usually regarded as safe enough to handle but could well be hazardous if inhaled and retained in the lungs for long enough.

If you combine the above facts with, for instance, sawing or grinding resins (particularly epoxies like Spurr's) then you might actually be increasing your exposure to the most hazardous component of the resin dust. Generally I try to avoid sawing and only ever trim blocks with razor blades or glass knives because the pieces are big and even if you could inhale some I doubt they'd ever reach your lungs.

Some of this is assimilated information from years ago, so I can't give you a particular source or reference. I suspect that you could easily check if there is a problem by putting a 'sticky' stub near to the exhaust of your vacuum cleaner and see if it picks up much. I've never tried it because I just simply avoid fine dust now.

Malcolm

Malcolm Haswell
e.m. unit
University of Sunderland
UK
e-mail: malcolm.haswell-at-sunderland.ac.uk



----- Original Message -----
} From: "Tindall, Randy D." {TindallR-at-missouri.edu}

The International Metallographic Society is seeking a volunteer to serve as
the Local Chair for the 2005 International Metallographic Contest which will
be held in conjunction with the 38th Annual Convention of the IMS and the
Microscopy and Microanalysis 2005 meeting in Honolulu, Hawaii in July of
2005. Interested individuals from the Honolulu area should contact the
Contest Chair, Jeff Stewart, at jeff-at-metallography.com for additional
information. We would like to have the position filled within the next few
weeks.

Thanks,

Jeff Stewart
International Metallographic Contest Chair
Metallographic Laboratory Manager
Stern-Leach Co.
49 Pearl Street
Attleboro, MA 02703 USA
Phone: 508-222-7400 extension 1329
Fax: 508-699-4030
E-mail: jeff-at-metallography.com



From MicroscopyL-request-at-ns.microscopy.com Thu May 6 08:43:47 2004



From: Philip Oshel :      peoshel-at-wisc.edu
Date: Thu, 06 May 2004 09:09:32 -0500
Subject: [Microscopy] Any used but good Gatan 626 cryostages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Good points, Malcolm. I forgot to mention that this thing comes with a
HEPA filter. Anyway, most of our resin fluff comes from block trimming
with razor blades and glass knives.

Randy

-----Original Message-----
} From: Malcolm Haswell [mailto:malcolm.haswell-at-sunderland.ac.uk]
Sent: Thursday, May 06, 2004 8:42 AM
To: MSA microscopy listserver
Cc: Tindall, Randy D.

Micromavens,

I'm hoping to find someone with a Gatan model 626
cryostage/cryoholder who wishes to sell it or donate it to a
university. Cryostation is not necessary, we've got that. Anyone out
there?
Thanks.

Phil
--
Philip Oshel
Supervisor, BBPIC microscopy facility
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706 - 1284
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)


From MicroscopyL-request-at-ns.microscopy.com Thu May 6 10:12:47 2004



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Thu, 6 May 2004 12:09:20 -0400
Subject: [Microscopy] marking tissues before wax embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

One of faculty here will be processing lengths of rat femoral artery
for light microscopy and needs to keep track of the distal vs
proximal ends. She tried marking one end with India Ink, but it was
washed out sometime during the dehydration/xylene steps.
Any ideas?
Thanks,
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Thu May 6 10:49:32 2004



From: Michael Dunlap :      mrdunlap-at-ucdavis.edu
Date: Thu, 6 May 2004 08:51:51 -0700
Subject: [Microscopy] Fixation of animal lung tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Beverly,

I was on a TEM project many years ago working with sheep and chicken
lungs. Jeff Weidner was the PI on the project. The problem in the
we had was to get the air out of the lungs before we could any of the
specimen prep procedures to work well. To many air bubbles and poor
infiltration.

We ended up use a standard EM procedure with each step done under low vacuum.

Mike

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
===================================================================================
Michael Dunlap
office (530) 752-0284
University of California lab (530) 752-5489
Chemical Engineering & Material Science Fax (530) 752-9554
110A Kemper Hall mrdunlap-at-ucdavis.edu
One Shields Ave.
http://www.chms.ucdavis.edu/
Davis CA, 95616
===================================================================================


From MicroscopyL-request-at-ns.microscopy.com Thu May 6 11:43:58 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Thu, 6 May 2004 09:51:48 -0700
Subject: [Microscopy] Re: marking tissues before wax embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 6, 2004, at 9:09 AM, Leona Cohen-Gould wrote:

} One of faculty here will be processing lengths of rat femoral artery
} for light microscopy and needs to keep track of the distal vs proximal
} ends. She tried marking one end with India Ink, but it was washed out
} sometime during the dehydration/xylene steps.
} Any ideas?
}
Dear Lee,
Could one end be cut straight across and the other cut at an angle? I
don't know how readily this could be done or whether it would be easy
to distinguish the ends, but if one cannot determine the shape of the
end, I'd wonder whether other morphology would be reliable.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Thu May 6 13:36:27 2004



From: Judy Murphy :      jmurphy-at-deltacollege.org
Date: Thu, 06 May 2004 11:22:25 -0700
Subject: [Microscopy] Re: Fixation of animal lung tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Beverly,
We have the same experience as Mike with both lung, brain (and a few
others) and most plant tissues. Before microwaving, we used the
standard protocols (G, followed by Os, etc.) and did all steps in a low
vacuum. You can use an evaporator using only the rough pump (Good idea
to protect where the vials are set so resin doesn't get on evaporator
plate_, if you don't have another vac chamber such as a vacuum oven.
Now that we microwave most of our tissues, we use the cold plate and
vacuum chamber for hard to penetrate tissues on all steps. It is likely
that the alcohols don't need to be done in the MW however it is faster.

NOTE about pulling the vacuum. One needs to WATCH when pumping the
samples. We pump on the samples and wait for the bubbles to rise. Just
before they get too close to the top, we turn off the vacuum, let the
bubbles settle a bit and start again. Eventually no bubbles appear.
This of course is likely more crucial in the resin steps since resin all
over everything is not nice! After the bubbles have subsided and don't
appear in quantity, we let the vial sit in the vacuum chamber under
vacuum but with the pump off, for the desired time of the step.

When using a microwave with a vacuum chamber it is important to note
WHERE the gauge actually is since if it is directly off the pump, then
when it pulls down to 20mmHg that is only at the pump port, not the
chamber, so that must be taken into consideration. We usually use that
approximate vacuum for the vac chamber on the microwave for most
fixations that are done in the microwave. It is hoped, sincerely hoped,
that eventually the manufacturers will build in a vacuum chamber with a
cold plate in the microwave so it isn't quite so messy to set up.

Good luck,
Judy Murphy

Michael Dunlap wrote:
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Hi Beverly,
}
} I was on a TEM project many years ago working with sheep and chicken
} lungs. Jeff Weidner was the PI on the project. The problem in the
} we had was to get the air out of the lungs before we could any of the
} specimen prep procedures to work well. To many air bubbles and poor
} infiltration.
}
} We ended up use a standard EM procedure with each step done under low vacuum.
}
} Mike
}
} } ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} } http://www.msa.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -------------------------------------------------------------------------------
} }
} } Below is the result of your feedback form (NJZFM-ultra-55). It was
} } submitted by (bwareham-at-utah.gov) from
} } http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html
} } on Wednesday, May 5, 2004 at 15:08:22
} } ---------------------------------------------------------------------------
} }
} } Email: bwareham-at-utah.gov
} } Name: Beverly Wareham
} }
} } Organization: Utah Veterinary Diagnostic Lab
} }
} } Title-Subject: [Microscopy] [Filtered] Fixation of animal lung tissue
} }
} } Question: Hi all,
} } Does anyone have a protocol for fixing animal lung tissue (sheep)
} } for EM and is willing to share it? I'd really appreciate it.
} }
} } Thanks,
} } Beverly Wareham
} } Utah Veterinary Diagnostic Laboratory
} } 435-797-1799
} } bwareham-at-utah.gov
} }
} } ---------------------------------------------------------------------------
}
} --
} ===================================================================================
} Michael Dunlap
} office (530) 752-0284
} University of California lab (530) 752-5489
} Chemical Engineering & Material Science Fax (530) 752-9554
} 110A Kemper Hall mrdunlap-at-ucdavis.edu
} One Shields Ave.
} http://www.chms.ucdavis.edu/
} Davis CA, 95616
} ===================================================================================


From MicroscopyL-request-at-ns.microscopy.com Thu May 6 14:26:10 2004



From: Bentley, Karen :      Karen_Jensen-at-URMC.Rochester.edu
Date: Thu, 6 May 2004 15:29:18 -0400
Subject: [Microscopy] Marking with India Ink

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Leona:

When you use India Ink it needs to be fixed to the tissue's surface you are
labeling with a picric acid solution such as Bouin's fixative.

Make the ink mark with the opposite end of a swab and then switch to the
cotton tipped end of the swab. Dipp into the Bouin's and dab the area
marked with the India Ink. Blot slightly dry with a Kim Wipe and put back
into formalin. It will be permanently fixed in place and can be processed
and embedded in paraffin.

Karen

Karen L. Bentley, M.S.(previously Jensen)
Associate Scientist & Manager
Electron Microscope Research Core
University of Rochester Medical Center
Rochester, NY 14642
585-275-1954





From MicroscopyL-request-at-ns.microscopy.com Thu May 6 14:53:29 2004



From: sbs38-at-email.byu.edu (by way of Ask-A-Microscopist)
Date: Thu, 6 May 2004 15:03:33 -0500
Subject: [Microscopy] AskAMicroscopist: protocols for viewing HSV-1

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (sbs38-at-email.byu.edu) from
http://www.msa.microscopy.org/Ask-A-Microscopist.html on Thursday,
May 6, 2004 at 10:14:58
---------------------------------------------------------------------------

Email: sbs38-at-email.byu.edu
Name: Briant Stringham

Organization: BYU

Education: Undergraduate College

Location: Provo, Utah, US

Question: I am collecting protocols for viewing HSV-1 through
negative staining, and I was hoping you could send what you feel is
the best protocol for viewing membrane and capsid integrity. Next, I
need to know the best stain and method for viewing empty or full
capsids.
Thank you for your reply

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu May 6 16:08:42 2004



From: Steven Cogswell :      cogswell-at-nbnet.nb.ca
Date: Thu, 06 May 2004 18:11:39 -0300
Subject: [Microscopy] [uProbe] Seeking Jeol JXA-733 Superprobe Parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello MSA folks;

We here at the University of New Brunswick have a Jeol JXA-733
Superprobe that has served us faithfully for many years. Sadly we have
had a catastrophic failure of the oil-cooled objective lens/pole piece
assembly, and so require a replacement. If you have working one and
would be willing to part with it, please contact us.

Thanks for your time,


Steven Cogswell, P.Eng.
UNB Microscopy and Microanalysis Facility
Fredericton, New Brunswick Canada
Phone (506) 453-4887





From MicroscopyL-request-at-ns.microscopy.com Thu May 6 18:32:09 2004



From: sara goldston :      cogold13-at-yahoo.com
Date: Thu, 6 May 2004 16:34:29 -0700 (PDT)
Subject: [Microscopy] Re: Marking with India Ink

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

You could try OsO4. Its guaranteed to darken on the
specimen and it won't wash out. The only concern is
how useful that end will be once its been treated with
Osmium. I know Osmium is used to fix fatty specimens
in order to give them more contrast. I would use a 1%
or maybe a 2% solution. It usually takes effect
within a few minutes, so you could maybe dip the
distal or promixal portion in the solution. Good
luck.
Sara Goldston
"Bentley, Karen" {Karen_Jensen-at-URMC.Rochester.edu}
wrote:

------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy
Society of America

Hi Leona:

When you use India Ink it needs to be fixed to the
tissue's surface you are
labeling with a picric acid solution such as Bouin's
fixative.

Make the ink mark with the opposite end of a swab and
then switch to the
cotton tipped end of the swab. Dipp into the Bouin's
and dab the area
marked with the India Ink. Blot slightly dry with a
Kim Wipe and put back
into formalin. It will be permanently fixed in place
and can be processed
and embedded in paraffin.

Karen

Karen L. Bentley, M.S.(previously Jensen)
Associate Scientist & Manager
Electron Microscope Research Core
University of Rochester Medical Center
Rochester, NY 14642
585-275-1954









__________________________________
Do you Yahoo!?
Win a $20,000 Career Makeover at Yahoo! HotJobs
http://hotjobs.sweepstakes.yahoo.com/careermakeover


From MicroscopyL-request-at-ns.microscopy.com Thu May 6 21:04:00 2004



From: David Elliott :      David.Elliott-at-yale.edu
Date: Thu, 6 May 2004 22:07:09 -0400
Subject: [Microscopy] American Laser

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Listers
I am looking for contact information for American Laser. Are they
still out there? How can I contact them?
Thanx
David



____________________

David Elliott Ph.D.

Yale University School of Medicine
Campus Address: TAC S160
333 Cedar Street
PO Box 208022
New Haven, CT   06520-8022

Tel: (203) 785-7572
Fax: (203) 785-6815



From MicroscopyL-request-at-ns.microscopy.com Fri May 7 13:26:32 2004



From: Steve Barlow :      sbarlow-at-sciences.sdsu.edu
Date: Fri, 7 May 2004 11:32:00 -0700
Subject: [Microscopy] Zeiss AXIOTRON

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear listers

I recently received the parts to a Zeiss Axiotron microscope, which
is no longer supported by the manufacturer. Does anyone have a
manual or schematics for this unit--I hate to waste the box of very
high quality lenses that came with it, and our engineering department
is very interested in getting the unit assembled and working.

Anyone that can help me out with this, please contact me.

thanks in advance

Steve Barlow
--
Dr. Steven Barlow
EM Facility/Biology Dept.
San Diego State University
5500 Campanile Drive
San Diego CA 92182-4614
phone: (619) 594-4523
fax: (619) 594-5676
email: sbarlow-at-sunstroke.sdsu.edu
http://www.sci.sdsu.edu/emfacility

Chairman, Educational Outreach subcommittee
promoting microscopy instruction and increased access to microscopes
Microscopy Society of America
http://www.msa.microscopy.com/


From MicroscopyL-request-at-ns.microscopy.com Fri May 7 14:34:41 2004



From: Frank.Karl-at-degussa.com
Date: Fri, 7 May 2004 15:37:29 -0400
Subject: [Microscopy] help: Image analysis training

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I'm hoping the microscopy community can help me. I had planned on taking a
course on imaging analysis and photomicrography at McCrone Research but it
was canceled (I was the only student.)

I have a real need to improve my imaging and image measurement skill. I
wish to do particle sizing on images captured via the SEM/TEM/light
microscope.

Does anyone know of short courses and such for the tyro? I would like take
one as soon as possible.

Thanks in advance!!!

Frank Karl
330-668-2235 Ext. 238




From MicroscopyL-request-at-ns.microscopy.com Fri May 7 18:15:57 2004



From: e.gagnepain-at-wanadoo.fr (by way of Ask-A-Microscopist)
Date: Fri, 7 May 2004 18:26:03 -0500
Subject: [Microscopy] AskAMicroscopist: FeAL for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (e.gagnepain-at-wanadoo.fr) from
http://www.msa.microscopy.org/Ask-A-Microscopist.html on Friday, May
7, 2004 at 10:04:31
---------------------------------------------------------------------------

Email: e.gagnepain-at-wanadoo.fr
Name: Gagnepain

Organization: ULP of Strasbourg

Education: Graduate College

Location: Strasbourg, France

Question: I have some samples of iron-aluminide (FeAL)and the
thickness is 2 millimeter. How to prepare thin layer of FeAl for an
observation with TEM (Transmission Electron Microscope) without
introduce defaults?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri May 7 18:28:32 2004



From: cryanez-at-criba.edu.ar (by way of MicroscopyListserver)
Date: Fri, 7 May 2004 18:38:18 -0500
Subject: [Microscopy] viaWWW: stain a polymer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form
(NJZFM-ultra-55). It was submitted by
(cryanez-at-criba.edu.ar) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html
on Friday, May 7, 2004 at 14:09:42
---------------------------------------------------------------------------

Email: cryanez-at-criba.edu.ar
Name: Julia YaÒez

Organization: CONICET

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I need to stain a polymer sample with
RuO4, by oxidation of RuCl3 with
sodium hypochlorite.
I don¥t know the appropiate reactives concentrations.
Could you help me about this procedure?.
Thanks in advance

Julia Y·Òez

Ing. MarÌa Julia Y·Òez TE: 54-291-486-1666 int.: 362
Lab.Microscopia ElectrÛnica FAX: 54-291-486-1527
CRIBABB e-mail: cryanez-at-criba.edu.ar
Camino La Carrindanga Km 7
(8000) BahÌa Blanca



---------------------------------------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Sat May 8 04:20:02 2004



From: Larry Stoter :      larry-at-cymru.freewire.co.uk
Date: Sat, 8 May 2004 09:53:56 +0100
Subject: [Microscopy] Re: AskAMicroscopist: FeAL for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I have never prepared specimens from this material, so I'm not
familiar with its properties.

Is it electrically conductive?

If so, to begin with I would spark errode with a trepanning tool to
get 3 mm dia. discs. Then mechanically grind and polish, preferably
with a dimple grinder, to ~100 um. Finally, electropolish, preferably
with a jet polisher.

Hirsch, Howie, Nicholson, Pashley and Whelan suggest, for Fe-Al
alloys acetic acid (133 cm3), water (7 cm3), chromic acid (25 g) with
a stainless steel cathode and rinses in acetic acid/methanol. Make
sure it remains cool ( {30 C). Voltage in the 25 V to 30 V range.

Possible problems:

1. Different phases may thin at different rates during
electropolishing - you'll need to change the polishing solution.

2. If it is very soft, the mechanical grinding and polishing may
create a deep damage layer - stop mechanical polishing sooner and use
electropolish earlier.

3. If there is stress in the material, thinning may cause a release
of stress and damage.

4. Safety officers

If you can't get electropolishing to work, ion milling should.
--
Larry Stoter
PLEASE NOTE
1. Any mail other than plain text will be automatically deleted.
2. Any mail, legitimate or not, apparently or actually from hotmail
will automatically be deleted.
:-)
PS - I'm an employee of JEOL UK.


From MicroscopyL-request-at-ns.microscopy.com Sat May 8 06:25:30 2004



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Sun, 09 May 2004 15:35:50 -0500
Subject: [Microscopy] Re: AskAMicroscopist: protocols for viewing HSV-1

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

This humble amateur histologist always used "alcoholic Bouin"
("Duboscq-Brazil Fluid") combined with gentle vacuum applied the way
Judy Murphy describes it, to fix small pieces of rabbit lung...

Try google with "alcoholic bouin" as the search string.


"alcoholic Bouin" (aka:"Duboscq-Brazil Fluid")

Stock Alcoholic Bouin's Solution:
80% ETOH 750.0 ml
formaldehyde 300.0 ml
Picric acid 5.0 gm

Mix solution well, stable 1 year.

Working Solution:
Stock solution 70.0 ml 14.0 ml
Acetic acid 5.0 ml 1.0ml
Add the acetic acid right before use.


Y.

-----Original Message-----
} From: Judy Murphy [mailto:jmurphy-at-deltacollege.org]
Sent: donderdag 6 mei 2004 20:22
To: Michael Dunlap
Cc: Microscopy-at-msa.microscopy.com; bwareham-at-utah.gov

briant

the question really is what is the source of specimen? are you looking
for clinical isolates or at culture material. if culture material, has
it been purified in any way. these things may all affect the
preparation and ultimate quality.

the first issue you have to address is grids. you will need grids with
some form of plastic film, be it collodion, formvar or parlodian. if
you are not set up to make grids, they can be obtained from most EM
supply houses. for short term situations, if you are not going to need
a lot of grids and have no one trained in their preparation, buying them
will be best for your use. you may also wish to carbon coat the grids
to stabilize the film. again, carbon coated plastic coated grids may be
purchased. if cost is a problem, i could probably manage to liberate
50-100 grids.

carbon coated grids introduce several problems, mainly due to the fact
that they are quite hydrophobic. however, they do become more
hydrophilic over time. also, you can treat the grids by glow discharge
(within 2-3 weeks of use) or treat them with alcian blue. usually
clinical material has so much serum or fecal protein present that if you
leave the sample on the grid for several minutes you overcome all
problems of hydrophobicity. we view 2500-3000 clinical samples, and
over 4000 total negative stain preparations/ year. at the risk of being
labeled a reprobate and cretin, we usually do not carbon stabilize the
grids used to examine most samples. the added stability is not enough
benefit for clinical material. if it is a sample of importance we may
use carbon coated grids. if so, we use alcian blue, or old grids.
either way, with our concentration methods, hydrophobicity is not a
problem.

you may wish to consider pretreatment of the sample to stabilize the
virion. we either add glutaraldehyde to a concentration of 0.1%, or
formaldehyde to a concentration of 1-2%. this allows us to neutralize
infectivity, maintain antigenicity and reactivity for IEM, and
stabilizes structure against stresses which may arise during
preparation. the samples need to be vortexed very well immediately to
limit clumping which may be associated with fixative mediated
crosslinking of virions. because of the shorter chain size and mono
aldehyde nature, some argue that formaldehyde is less likely to result
in crosslinking, and is better than glutaraldehyde. personally, we have
never seen any difference between the two.

you do not mention mounting the sample on the grid. i assume you have
some method of preference. if not, please ask, we all have our
favourite methods and are more than willing to tell you them.

as far as staining, most viral electron microscopists have their own
preferences. that does not make any one method 'best', however. we use
several stains routinely, and do a panel 14 different negative stains
whenever we start a new project to confirm the best set of conditions.
that said, the two major staining methods use tungsten or uranyl salts.

the most common tungsten salt is 1.5 to 2% phosphotungstic acid (PTA).
to eqalize the different methods, i have adjusted the concentrations of
stains in my labs to maintain an equal amount of heavy metal atoms
available in the staining solution. therefore, we use 2.5mM PTA. this
works out to be approximately 1.6%. the pH is adjusted to 7.0. the
best method for adjusting pH with this stain is with NaOH. early
studies by Horne and his associates implicated the loss of membrane
integrity with pH adjustment by KOH. if you were working with paramyxo
or myxo, and wanted to release the nucleocapsid, this would be a good
method. for the greater part you do not wish to disrupt the membranes.
with most herpes preparations, you will already see a lot of naked
complete and empty nucleocapsids as it is. we also add 25 micrograms/ml
bacitracin as a surfactant. this helps with stain spreadability and
penetration. other small molecules are available for use. the most
common is probably BSA. again, the addition of fixative will also
stabilize the virions against loss of membranes due to stain
characteristics.

we use two different versions of uranyl stains, uranyl acetate(UA) and
uranyl sulfate (US). both of these must be used at low pH. in fact, if
you try to adjust the pH to around 4.0 you start to get problems keeping
material in solution - especially if your pH meter is not as accurate as
you think. as a rule, we use the material at pH of about 3.5. again,
most protocols call for 2% solutions. adjusting to molar concentrations
and maintaining equivalent heavy metal atoms, we use a 60mM solution.
this represents a 2.5% UA or 2.2% US stain. if you must work with
higher pH, it has been suggested that 1% UA can be adjusted to pH values
as high as 7.0 by the addition of 0.1% ammonium acetate and 0.5% EDTA
(disodium salt). i have not tried this, so i do not know how well it
works. rumour is that it is quite grainy and sublimates badly.

each stain system has its own properties. many feel particles degrade
in PTA but not in UA. PTA has a higher anhydrous density, so you see
more contrast, which could be bad or good. UA tends to be more granular
when imaging, while PTA is low granularity. here the issue is purely
what you think would 'look best'. UA/US tends to sublimate in the beam,
leading to a perceived column contamination problem, while PTA is very
stable, leading some to argue that the column is less contaminated by
vaporizing stain.

the best suggestion - try both PTA and a uranyl stain. i would also
refer you to a recent review: Hazelton, P.R. and Gelderblom, H.R. (2003)
The use of the electron microscope for rapid diagnosis of viral agents
in emergent situations. Emerging Infectious Diseases 9:294-303. it will
give some additional methods and general information. yes, i'm human,
it is one of my articles, but i think it's good, anyway.

hope these thoughts help. let me know if you want specific protocols.
we have them all in electronic file format, albeit WordPerfect files, so
they can be sent easily.

paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-954
Cell:204-781-1502
Fax:204-789-3926


From MicroscopyL-request-at-ns.microscopy.com Sun May 9 19:06:16 2004



From: MICROCOBB-at-aol.com
Date: Sun, 9 May 2004 20:10:41 EDT
Subject: [Microscopy] Re: RE: Photomicrography and the digital age

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html




From MicroscopyL-request-at-ns.microscopy.com Mon May 10 00:27:40 2004



From: hunny_pot-at-optusnet.com.au (by way of Ask-A-Microscopist)
Date: Mon, 10 May 2004 00:34:14 -0500
Subject: [Microscopy] AskAMicroscopist: shrinkage due to fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (hunny_pot-at-optusnet.com.au) from http://www.msa.microscopy.com/Ask-A-Microscopist.html on Sunday, May 9, 2004 at 21:45:15
---------------------------------------------------------------------------

Email: hunny_pot-at-optusnet.com.au
Name: Linda

Organization: The University of Sydney

Education: Undergraduate College

Location: Australia

Question: Hello:

I am doing research paper on quanification of shrinkage due to fixation. There are some questions that I like answers to:

1. what are some of the methods used to measure shrinkage?

2. how the effects of foxation can be differentiated from other possible sources of shrinkage?

3. the practices and clincal signifance of quantifying shrinkage due to fixation.

I have been looking around, but I can't find my answers. Could you please help me? Thanks

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon May 10 07:57:18 2004



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Mon, 10 May 2004 15:01:05 +0200
Subject: [Microscopy] PEELS wanted

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi

We are looking for a second hand Gatan 666 PEELS, to fit on VG HB5 STEM.
Perheps are there somme which are sleeping in cupboards ! It doesn't
matter if modifications are necessary. Preferently, for administrative
raions, we would prefer a europeen source, but all propositions are
welcome.

Answer directly to my collegue, at
Daniel.Spor-at-ipcms.u-strasbg.fr

Many Thanks.

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr




From MicroscopyL-request-at-ns.microscopy.com Mon May 10 08:04:46 2004



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Mon, 10 May 2004 15:08:33 +0200
Subject: [Microscopy] EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi

What type of silver charged epoxy will be used to make embedding of powder
as standards for EDS microanalysis. If one looks at Epotek's catalog,
there are so much type of epoxys, with very different properties, shelf
live and price. So what is the best compromise. If someone has an advice.

Many thanks

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr




From MicroscopyL-request-at-ns.microscopy.com Mon May 10 08:53:24 2004



From: Geoff McAuliffe :      mcauliff-at-umdnj.edu
Date: Mon, 10 May 2004 09:57:11 -0700
Subject: [Microscopy] Re: AskAMicroscopist: shrinkage due to fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Linda:

A great deal of research was done on these questions in the late
1940's-1960's and you won't find it on the internet. For starters, get a
hold of "Principles of Biological Microtechnique" by John R. Baker,
published in 1958. He discusses various fixatives, their penetration
rates, how they do or do not promote shrinkage (in the fix or in
subsequent processing), etc. I will try to dig up some references to
specific papers later today.

Geoff

by way of Ask-A-Microscopist wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************





From MicroscopyL-request-at-ns.microscopy.com Mon May 10 09:43:36 2004



From: gary.m.brown-at-exxonmobil.com
Date: Mon, 10 May 2004 09:49:00 -0500
Subject: [Microscopy] Re: viaWWW: stain a polymer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


"The statements and opinions expressed here by Gary M. Brown represent
neither those of ExxonMobil Corporation nor its affiliates."

Julia,

Prepare only the amount of stain that you will use at that moment. For 1
ml of RuO4 stain, use 0.02 g RuCl(subscript: 3 )(superscript: .) nH
(subscript: 2)O [14898-67-0] and 1 ml of NaOCl [7681-52-9].

I recommend that you purchase fresh hypochlorite every few months since the
concentration of hypochlorite in solution does diminish with time. I
discourage using bleach for the same reason, i.e. one is never sure how
long the bottle of bleach has been on the shelf. Keep the RuCl(subscript:
3) hydrate in a nitrogen desiccator when not in use.

The original reference on the use of the in-situ preparation of RuO4 is:
-- Montezinos, D., Well, B.G. and Burns, J.L., J. Polymer Sci.: Polym.
Lett. Ed., 1985, 23, 421.

Another important reference pertains to vapor-staining with RuO4 (using
RuO4 crystals, which are very difficult to find):
-- Sano, H., Usame, T. and Nakagawa, H., Polymer, 1986 27, 1497.

I hybridized the methods of Sano and Montezinos, i.e. using vapor staining
above the in-situ preparation of RuO4, and invariably get outstanding
results:
-- Brown, G. M. and Butler, J. H., Polymer, 1997, 38 (15), 3937.

Although the Montezinos and Sano papers are key, their practicality is
limited. Montezinos recommends staining in the solution using precursors
that are dirt cheap and readily available. Sano convinced me, however,
that solution staining is not a good practice because it produces severe
and uncontrolled oxidation of the outer surface of the sample. Sano's
vapor-staining method is beautiful but the solid RuO4 crystals he uses are
virtually impossible to find in the United States and are very expensive.
I simply combined the two methods and worked out staining durations. Refer
to the appendix of my paper for details including safe use, handling and
disposal as well as sample preparation from beginning to end and staining
durations for some polymer systems.

Happy staining,

Gary M. Brown
ExxonMobil Chemical Company
Baytown Technology & Engineering - West
5200 Bayway Drive
Baytown, Texas 77520-2101
phone: (281) 834-2387
fax: (281) 834-2395
e-mail: Gary.M.Brown-at-ExxonMobil.com



cryanez-at-criba.edu.ar
(by way of To: microscopy-at-ns.microscopy.com
MicroscopyListserver) cc:
Subject: [Microscopy] viaWWW: stain a polymer

05/07/04 06:38 PM







------------------------------------------------------------------------------

The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Below is the result of your feedback form
(NJZFM-ultra-55). It was submitted by
(cryanez-at-criba.edu.ar) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html
on Friday, May 7, 2004 at 14:09:42
---------------------------------------------------------------------------

Email: cryanez-at-criba.edu.ar
Name: Julia YaÒez

Organization: CONICET

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I need to stain a polymer sample with
RuO4, by oxidation of RuCl3 with
sodium hypochlorite.
I don¥t know the appropiate reactives concentrations.
Could you help me about this procedure?.
Thanks in advance

Julia Y·Òez

Ing. MarÌa Julia Y·Òez TE: 54-291-486-1666 int.: 362
Lab.Microscopia ElectrÛnica FAX: 54-291-486-1527
CRIBABB e-mail: cryanez-at-criba.edu.ar
Camino La Carrindanga Km 7
(8000) BahÌa Blanca



---------------------------------------------------------------------------









From MicroscopyL-request-at-ns.microscopy.com Mon May 10 09:45:04 2004



From: Joel Sheffield :      jbs-at-temple.edu
Date: Mon, 10 May 2004 10:50:40 -0400
Subject: [Microscopy] Re: AskAMicroscopist: shrinkage due to fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

You may be interested in a paper I published a few years ago in which
I measured the size distribution of cells in suspension after
different fixation conditions. Develop. Biol 23:36-61 (1970)

Joel




}
}
} ----------------------------------------------------------------------
} -------- The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} ----------------------------------------------------------------------
} ---------
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (hunny_pot-at-optusnet.com.au) from
} http://www.msa.microscopy.com/Ask-A-Microscopist.html on Sunday, May
} 9, 2004 at 21:45:15
} ----------------------------------------------------------------------
} -----
}
} Email: hunny_pot-at-optusnet.com.au
} Name: Linda
}
} Organization: The University of Sydney
}
} Education: Undergraduate College
}
} Location: Australia
}
} Question: Hello:
}
} I am doing research paper on quanification of shrinkage due to
} fixation. There are some questions that I like answers to:
}
} 1. what are some of the methods used to measure shrinkage?
}
} 2. how the effects of foxation can be differentiated from other
} possible sources of shrinkage?
}
} 3. the practices and clincal signifance of quantifying shrinkage due
} to fixation.
}
} I have been looking around, but I can't find my answers. Could you
} please help me? Thanks
}
} ----------------------------------------------------------------------
} -----
}


Joel B. Sheffield, Ph.D.
Biology Department, Temple University
1900 North 12th Street
Philadelphia, PA 19122
jbs-at-temple.edu
(215) 204 8839, fax (215) 204 0486
http://astro.temple.edu/~jbs



From MicroscopyL-request-at-ns.microscopy.com Mon May 10 09:49:51 2004



From: Tomic, Peter \(Peter\) :      ptomic-at-agere.com
Date: Mon, 10 May 2004 10:55:26 -0400
Subject: [Microscopy] EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Jacques;

What is the material of interest? You need to say. Obviously it can't be silver but you may want to check to see if you have silver spectrally overlapping the material or element of interest. Is your concern primarily sample charging therefore you want a conductive epoxy?

Regards,

Peter Tomic
Agere Systems

-----Original Message-----
} From: Faerber Jacques [mailto:Jacques.Faerber-at-ipcms.u-strasbg.fr]
Sent: Monday, May 10, 2004 9:09 AM
To: Microscopy Society of America


Hi

What type of silver charged epoxy will be used to make embedding of powder
as standards for EDS microanalysis. If one looks at Epotek's catalog,
there are so much type of epoxys, with very different properties, shelf
live and price. So what is the best compromise. If someone has an advice.

Many thanks

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr






From MicroscopyL-request-at-ns.microscopy.com Mon May 10 10:06:37 2004



From: gary.m.brown-at-exxonmobil.com
Date: Mon, 10 May 2004 10:12:19 -0500
Subject: [Microscopy] Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


"The statements and opinions expressed here by Gary M. Brown represent
neither those of ExxonMobil Corporation nor its affiliates."

Jacques,

I spent a considerable amount of time working on conductive embedding
media. See my abstract in the 2004 M&M Proceedings.

The problem with virtually all commercial electrically conductive embedding
media that I have seen is that they offer only macroscopic conductivity.
This means that expanses of unfilled, non-conducting and highly charging
epoxy will be found throughout the sample. We need microscopic
conductivity in which the embedding medium is highly conductive and does
not contain non-conductive domains of epoxy. For expoxy-embedding
applications, I recommend compounding (mixing) carbon black filler into the
resin of choice (without catalyst or accelerator) in a mortar and pestle.
Use equal parts by weight of the filler powder and epoxy. At the last
minute before embedding, mix the accelerator into the resin / carbon black
compound as quickly and thoroughly as possible. Embed the sample in this
thick paste.

If well mixed, the embedding medium will not charge at all. Also, since
only carbon and oxygen are present in any appreciable concentration, you
will not have problems with interference lines from metal filler particles,
i.e. silver, etc.

Have fun.

Gary M. Brown
ExxonMobil Chemical Company
Baytown Technology & Engineering - West
5200 Bayway Drive
Baytown, Texas 77520-2101
phone: (281) 834-2387
fax: (281) 834-2395
e-mail: Gary.M.Brown-at-ExxonMobil.com



Faerber Jacques
{Jacques.Faerber-at-ipcms.u- To: Microscopy Society of America {Microscopy-at-MSA.Microscopy.Com}
strasbg.fr} cc:
Subject: [Microscopy] EDS standards embedding

05/10/04 08:08 AM







------------------------------------------------------------------------------

The Microscopy ListServer -- Sponsor: The Microscopy Society of America



Hi

What type of silver charged epoxy will be used to make embedding of powder
as standards for EDS microanalysis. If one looks at Epotek's catalog,
there are so much type of epoxys, with very different properties, shelf
live and price. So what is the best compromise. If someone has an advice.

Many thanks

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr










From MicroscopyL-request-at-ns.microscopy.com Mon May 10 10:12:59 2004



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Mon, 10 May 2004 10:14:46 -0500
Subject: [Microscopy] Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Our standards from Tsousimis came prepared with regular epoxy and a coating
of carbon. We have used the same procedure for our homemade standards. Many
of the standard materials are not conductive anyway, so the conductivity
has to come from a coating. Of course a "conductive" epoxy should not hurt.

I presume you will be not be collecting spectra in variable pressure mode.
The scattering of the beam from the residual atmosphere would produce a
signal from the silver or nickel in your epoxy and would complicate things.

I also advise you to prepare the sample and standards the same. I thought I
was competent at EDS standards and realized fairly recently the difference
that can be caused by different thickness of C coatings (or lack thereof)
between standards and unknowns.

Warren Straszheim

At 08:08 AM 5/10/2004, you wrote:

} Hi
}
} What type of silver charged epoxy will be used to make embedding of powder
} as standards for EDS microanalysis. If one looks at Epotek's catalog,
} there are so much type of epoxys, with very different properties, shelf
} live and price. So what is the best compromise. If someone has an advice.
}
} Many thanks
}
} J. Faerber
} IPCMS-GSI
} (Institut de Physique et Chimie des Matériaux de Strasbourg
} Groupe Surface et Interfaces)
} 23, rue de Loess ; BP43
} 67034 Strasbourg CEDEX 2
} France
}
} Tel 00 33(0)3 88 10 71 01
} Fax 00 33(0)3 88 10 72 48
} E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr




From MicroscopyL-request-at-ns.microscopy.com Mon May 10 10:28:32 2004



From: Cheri Owen :      cowen-at-cmb.biosci.wayne.edu
Date: Mon, 10 May 2004 11:28:24 -0400
Subject: [Microscopy] Cell fraction negative staining protocol

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I want to find a simple TEM protocol for negative staining of microsomal
preparations. Does anyone have suggestions?

Thank you,
Cheri Owen
Cheri Owen, PhD
Dept. Emergency Medicine
Wayne State University
Detroit, Mi 48201


From MicroscopyL-request-at-ns.microscopy.com Mon May 10 10:56:54 2004



From: Eric Anderson :      andersone1-at-southernct.edu
Date: Mon, 10 May 2004 12:03:01 -0400
Subject: [Microscopy] EM400 Chiller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ooh, one more question:

The Philips EM400 installations I know of all utilized a Haskris
chiller. We're looking to buy a new chiller and they are considerably
cheaper from Neslab. Would anyone recommend against going with Neslab
for any reason?

Many thanks for any tips!
Best to all,
Eric Anderson
SCSU Physics Department
501 Crescent Street
New Haven, CT 06515
203-392-6468





From MicroscopyL-request-at-ns.microscopy.com Mon May 10 10:56:54 2004



From: Eric Anderson :      andersone1-at-southernct.edu
Date: Mon, 10 May 2004 12:02:33 -0400
Subject: [Microscopy] Camera C-mount on EM400

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi folks,

Does anyone know of an archived discussion or a web site anywhere that
describes adapting a C-mount for a digital camera to the old Philips
EM400 TEM? Perhaps by custom machining a cover blank for the 35mm
camera mounting location?

We recently purchased a Pax-it 2.1 Mpixel digital camera to use with a
new inverted metallurgical LM, but after receiving the camera, funding
for the microscope didn't materialize, so we're thinking of using the
Pax-it on our EM400.

Many thanks for any tips!
Best to all,
Eric Anderson
SCSU Physics Department
501 Crescent Street
New Haven, CT 06515
203-392-6468





From MicroscopyL-request-at-ns.microscopy.com Mon May 10 13:52:13 2004



From: Thomas, Frank :      FThomas-at-nrcan.gc.ca
Date: Mon, 10 May 2004 15:58:05 -0300
Subject: [Microscopy] EM400 Chiller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Eric -

The only thing I don't particularly care for in our (?8-year-old)
Neslab chiller is the fact that the water reservoir is pretty well sealed
up, like a miniature 45-gallon drum on its side, so you can't see what's
going on in there. The Haskris chillers always used to have a water tank
with a lid, making for much easier service and monitoring of water
condition. Other than that, both types are pretty well just pumps directly
driven by electric motors with some miscellaneous tubing and controls. Both
brands used to use the same pumps (made by Procon of Tennessee), and
probably still do.
Our Neslab has been pretty trouble-free; it's gone through one motor
so far and at least two pumps, but, as they say, your mileage may vary.

Frank

F.C. Thomas
FThomas-at-NRCan.gc.ca, 902-426-4635, facsimile 902-426-6152
GSC Atlantic
Natural Resources Canada, Bedford Institute of Oceanography,
P.O. Box 1006, Dartmouth, Nova Scotia B2Y 4A2
Ressources naturelles Canada, l'Institut Oceanographique du Bedford, B.P.
1006, Dartmouth, (Nouvelle-Ecosse)
B2Y 4A2
Government of Canada/Gouvernement du Canada


-----Original Message-----
} From: Eric Anderson [mailto:andersone1-at-southernct.edu]
Sent: Monday, May 10, 2004 1:03 PM
To: MSA listserver

Ooh, one more question:

The Philips EM400 installations I know of all utilized a Haskris
chiller. We're looking to buy a new chiller and they are considerably
cheaper from Neslab. Would anyone recommend against going with Neslab
for any reason?

Many thanks for any tips!
Best to all,
Eric Anderson
SCSU Physics Department
501 Crescent Street
New Haven, CT 06515
203-392-6468





From MicroscopyL-request-at-ns.microscopy.com Mon May 10 15:32:29 2004



From: Deb Gaynor :      dgaynor-at-analyticalservices.com
Date: Mon, 10 May 2004 16:36:28 -0400
Subject: [Microscopy] EM400 Chiller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We have been using a Neslab chiller since the beginning of time (I can't
find the original records). It runs without much complaint, most of the
time. We have replaced the motor (just an electric motor available from
Grainger's) once, replaced the pump once, and had the old pump rebuilt by
Procon once. We have replaced the shear pin ("nylon coupling")in the pump
head a number of times (maybe as much as once a year), and recharged the
compressor once in the last 6 years.

I've heard that the Haskris is a "better" chiller...but the Neslab seems to
do the job just fine. Their service support--by phone--is wonderful, and
they're in southern NH if you need to send something to them.

I highly recommend having a supply of shear pins available, and definitely
have them teach you how to change them if you never have done this
before--it's a little tricky the first time.

No interest in Neslab other than long experience as a user...

Deborah H. Gaynor, Ph.D.
Scanning Electron Microscopist
Analytical Services, Inc.
PO Box 515
Williston, VT 05495
802-878-5138 x21

-----Original Message-----
} From: Thomas, Frank [mailto:FThomas-at-nrcan.gc.ca]
Sent: Monday, May 10, 2004 2:58 PM
To: 'Eric Anderson'; MSA listserver

Eric -

The only thing I don't particularly care for in our (?8-year-old)
Neslab chiller is the fact that the water reservoir is pretty well sealed
up, like a miniature 45-gallon drum on its side, so you can't see what's
going on in there. The Haskris chillers always used to have a water tank
with a lid, making for much easier service and monitoring of water
condition. Other than that, both types are pretty well just pumps directly
driven by electric motors with some miscellaneous tubing and controls. Both
brands used to use the same pumps (made by Procon of Tennessee), and
probably still do.
Our Neslab has been pretty trouble-free; it's gone through one motor
so far and at least two pumps, but, as they say, your mileage may vary.

Frank

F.C. Thomas
FThomas-at-NRCan.gc.ca, 902-426-4635, facsimile 902-426-6152
GSC Atlantic
Natural Resources Canada, Bedford Institute of Oceanography,
P.O. Box 1006, Dartmouth, Nova Scotia B2Y 4A2
Ressources naturelles Canada, l'Institut Oceanographique du Bedford, B.P.
1006, Dartmouth, (Nouvelle-Ecosse)
B2Y 4A2
Government of Canada/Gouvernement du Canada


-----Original Message-----
} From: Eric Anderson [mailto:andersone1-at-southernct.edu]
Sent: Monday, May 10, 2004 1:03 PM
To: MSA listserver

Ooh, one more question:

The Philips EM400 installations I know of all utilized a Haskris
chiller. We're looking to buy a new chiller and they are considerably
cheaper from Neslab. Would anyone recommend against going with Neslab
for any reason?

Many thanks for any tips!
Best to all,
Eric Anderson
SCSU Physics Department
501 Crescent Street
New Haven, CT 06515
203-392-6468







From MicroscopyL-request-at-ns.microscopy.com Mon May 10 17:03:35 2004



From: David Knecht :      knecht-at-uconn.edu
Date: Mon, 10 May 2004 18:09:33 -0400
Subject: [Microscopy] gfp antibody

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Has anyone tried any of the anti-GFP antibodies on the market for
western blotting? Many seem to be cheaper than Clontech for the same
amount but we don't want to buy an untested antibody. Thanks- Dave
--

Dr. David Knecht
Department of Molecular and Cell Biology
University of Connecticut
91 N. Eagleville Rd. U-3125
Storrs, CT 06269-3125
knecht-at-uconn.edu
860-486-2200 860-486-4331 (fax)
home page: http://www.sp.uconn.edu/~mcbstaff/knecht/knecht.html



From MicroscopyL-request-at-ns.microscopy.com Mon May 10 18:18:11 2004



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 10 May 2004 16:23:59 -0700
Subject: [Microscopy] Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

What are you trying to look at with your EDS?
What kind of EDS is it?

The "normal" calibration of EDS is done at Al
K alpha (1.486KeV) and Cu K alpha (8.040KeV).
This sets resolution of the EDS according to these
two peaks. The other checks I do are done at C, O and F
and at N & O. And finally, FWMH at Mn.

For these measurements, I use X-Checker and X-Checker Extra.
For C, O and F, I use an Al stub with a Carbon sticky tab
with Teflon tape about 50% covering the stub and the other
50% split between sticky tab and bare Aluminum pin stub.
These readings do not calibrate the system. They produce
resolution figures for ability to resolve light elements.
You can do similar analysis for heavier elements with other
stub configurations.

The rationale for light element work is to see if the
EDS system, over time, drifts in the wrong direction.
Then I would know that something is going bad, either window
problems or EDS system problems. You will see this right
away at low Z.

If you are looking for ISO standards, I'm not sure how
that would work. Cu is Cu and Al is Al. The Cu and Al
peaks are used. If any foreign material was present, it
is simply ignored. The EDAX calibration routine only looks
at the Al and Cu peaks.

gary g.


At 06:08 AM 5/10/2004, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Mon May 10 18:27:40 2004



From: Rick A. Harris :      raharris-at-ucdavis.edu
Date: Mon, 10 May 2004 16:32:56 -0700
Subject: [Microscopy] RE: EM400 Chiller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have been running my Haskris, on my EM410, virtually continuously,
since November 1982. I have replaced no parts. I have another Haskris
on my 1999 Hitachi S3500-N that has an intermittent algae problem, and
has also needed the starter capacitor on the motor replaced.

Rick A. Harris, exDirector
Microscopy and Imaging Facility
Section of Molecular and Cellular Biology
1241 Life Sciences Addition
University of California
Davis, CA
530 752 2914
http://microscopy.mcb.ucdavis.edu
raharris-at-ucdavis.edu




From MicroscopyL-request-at-ns.microscopy.com Tue May 11 02:06:04 2004



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Tue, 11 May 2004 09:09:45 +0200
Subject: [Microscopy] Re: EDS standards embedding (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Some precisions, as Peter has asked :

The question comes for a job on thin film of materials such Sr2TiMoO4 and
CoFe2O4 grown by laser ablation. The goal is to control the stochioametric
variation from the films, in comparaison with the target material, which
we have as a powder, or as sintered target. So I need to embed the powder,
or pieces of the sintered bulk, to make a polished surface of
it. I reed in a paper from Geller in the NIST bull. mentionned by someone
on the list about the use of silver charged epoxy. As I use some by the
way for fixing sample for spark machining, I was in question about what
kind could do the job for my standards.

As I have understand the electrical conductivity in such materials will be
done by tunneling effect between the silver grain, through the epoxy. So
it is not so macroscopic. I'll try the methode with carbon black. Seems to
be interesting and much cheaper than all the commercial silver epoxys. And
with carbon, the problem of siver L lignes signal will be avoid.

These standards are used as references for thin films analysis, using
Stratagem thin film software.

Some other suggestions ?
Many thanks.


J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr

On Mon, 10 May 2004, Tomic, Peter (Peter) wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Jacques;
}
} What is the material of interest? You need to say. Obviously it can't be silver but you may want to check to see if you have silver spectrally overlapping the material or element of interest. Is your concern primarily sample charging therefore you want a conductive epoxy?
}
} Regards,
}
} Peter Tomic
} Agere Systems
}
} -----Original Message-----
} } From: Faerber Jacques [mailto:Jacques.Faerber-at-ipcms.u-strasbg.fr]
} Sent: Monday, May 10, 2004 9:09 AM
} To: Microscopy Society of America
} Subject: [Microscopy] EDS standards embedding
}
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
}
} Hi
}
} What type of silver charged epoxy will be used to make embedding of powder
} as standards for EDS microanalysis. If one looks at Epotek's catalog,
} there are so much type of epoxys, with very different properties, shelf
} live and price. So what is the best compromise. If someone has an advice.
}
} Many thanks
}
} J. Faerber
} IPCMS-GSI
} (Institut de Physique et Chimie des Matériaux de Strasbourg
} Groupe Surface et Interfaces)
} 23, rue de Loess ; BP43
} 67034 Strasbourg CEDEX 2
} France
}
} Tel 00 33(0)3 88 10 71 01
} Fax 00 33(0)3 88 10 72 48
} E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr
}
}
}
}
}





From MicroscopyL-request-at-ns.microscopy.com Tue May 11 03:24:48 2004



From: mivalmar-at-yahoo.es (by way of MicroscopyListserver)
Date: Tue, 11 May 2004 15:29:02 -0500
Subject: [Microscopy] viaWWW: immunohistochemistry of PPAR alpha

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Rick

this may be drifting off topic a bit, but we used to have algae problems with our water cooler until I covered the translucent plastic hose with aluminium foil. There doesn't seem to have been a problem for over a year and there is little need to clean out the system these days.

I know that you can get black pressure hoses as an alternative but you can't buy them from your local supermarket for ~ 1 UK pound and fit them in a few minutes while the system is running.

Malcolm

Malcolm Haswell
e.m. unit
University of Sunderland
UK

e-mail: malcolm.haswell-at-sunderland.ac.uk


----- Original Message -----
} From: "Rick A. Harris" {raharris-at-ucdavis.edu}

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (mivalmar-at-yahoo.es) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, May 11, 2004 at 15:10:15
---------------------------------------------------------------------------

Email: mivalmar-at-yahoo.es
Name: Miguel V.Serrano

Organization: C.Haya Hospital

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I want to ask if someone has tryed to do an immunohistochemistry of PPAR alpha (peroxisome proliferator receptor alpha) in rat liver or another tissue ??. Thanks.


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue May 11 15:21:51 2004



From: donovanl-at-ufl.edu (by way of MicroscopyListserver)
Date: Tue, 11 May 2004 15:28:23 -0500
Subject: [Microscopy] viaWWW: Nikon Labophot 1x objective

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (donovanl-at-ufl.edu) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, May 11, 2004 at 12:50:14
---------------------------------------------------------------------------

Email: donovanl-at-ufl.edu
Name: Laura Donovan

Organization: University of Florida College of Veterinary Medicine

Title-Subject: [Microscopy] [Filtered] MListserver:Nikon Labophot 1x objective

Question: I am looking for a 1x objective for a Nikon Labophot, and having a difficult time.
Lab phone: (352)392-4700 ext. 3845
ask for Rose or Laura.

Thank you.


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed May 12 00:36:05 2004



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Wed, 12 May 2004 17:45:19 +1200
Subject: [Microscopy] Evolved-Gas Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


I'm sorry that this is a bit of a deviation, but, knowing how diverse this list is, somebody
may be able to help:-

I have a Stanton-Redcroft Thermal Analyser, an STA-1600, the company in its last
years was acquired by Polymer Laboratories and then by Rheometrics, and I'm
interested in exploring the possibility of hooking analytical instruments, maybe NDIR or
even mass-spec, to the gas outlet line, to analyse the evolved gases (what was referred
to some time ago, so nicely, by Stanton-Redcroft as "hyphenated techniques").

Does anyone have such a setup?

Presumably one would use a heated PTFE interface tube, but the furnace internal
volume is fairly large, and not necessarily efficiently swept by the 50 ml/min or so of
purge gas, and what I'm wondering about is the time resolution of such EGA in this
instrument.

cheers

rtch

--
Ritchie Sims Ph D Phone : 64 9 3737599 ext 87713
Microanalyst Fax : 64 9 3737435
Department of Geology email : r.sims-at-auckland.ac.nz
The University of Auckland
Private Bag 92019
Auckland
New Zealand



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 04:52:33 2004



From: michael shaffer :      michael-at-shaffer.net
Date: Wed, 12 May 2004 07:27:50 -0230
Subject: [Microscopy] RE: Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Gary writes ...

} The "normal" calibration of EDS is done at Al
} K alpha (1.486KeV) and Cu K alpha (8.040KeV).
} ...

Out of curiosty, why would a calibration use these
2 peaks rather than CuK and CuL?? Understandably,
the L-line would be a convolution of the L-family,
but I would think the Al K line is in a more difficult
location relative to the slope in the continuum(???)
Is there any preference with respect to EDX window
type? (e.g., Be v UTW)

cheerios ... shAf :o)
Avalon Peninsula, Newfoundland
www.micro-investigations.com (in progress)



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 05:04:38 2004



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Wed, 12 May 2004 12:15:08 +0200
Subject: [Microscopy] EM masters need help. Please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear colleagues

We are looking for the RIGHT answers.
Addressing different institutions across the globe appear to be a very difficult.
We have been sent into circles for months without one clear cut answer.

Our EM technician has a HND in Environmental Science. He has been in the EM Unit for just over
a year and is slowly getting the hang of things. We would like to promote him up the ranks.
He has got the right aptitude, has matured and is responsible. Nice to find people like that still.

For him to be promotable, the University require at least a first degree (Senior Technician)
and Masters (Chief Technician). Without going through the rest of the gory details, he need further study. It is the policy of the University of Botswana to send citizens abroad rather than to study locally to enrich the nation academically.

The ranking of preference is 1) South Africa, 2) Australia, 3) USA, 4) UK.

He would prefer to do a EM focused masters. If he can do a pure Electron Microscopy Masters through course and lab work the better. The preference will be in the field of Material Science if he can not do a masters in "electron microscopy".

1) For him to be send by the University for a first degree and a masters consecutively, he must find at least partial sponsorship/bursary.
2) Who must he contact. (Please one name only in your institution, not a string of people!)
3) Can he get Credit for his current qualification?
4) Academic year start, requirements, and duration of study.

We are keeping our fingers crossed.
Thanks

Since my UB mail address is not reliable please send
a copy of all urgent mail to coetzeesh-at-yahoo.co.uk

Mr S. H. Coetzee
Electron Microscope Unit
University of Botswana
Private Bag 0022
Gaborone
Botswana

Phone : +267 355 2462/5222
Mobile : +267 714 55701 (Now active)
Fax : +267 318 5097
e-mail : coetzees-at-mopipi.ub.bw



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 06:52:59 2004



From: White, Woody N. :      nwwhite-at-bwxt.com
Date: Wed, 12 May 2004 06:56:28 -0500
Subject: [Microscopy] Evolved-Gas Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Ritchie,

FWIW, we have a Leybold quadrapole RGA which we often connect to our vacuum
furnace to monitor evolved gasses. Since the RGA operates at much higher
vacuum than the furnace, the interface is made using conflat flanged s-steel
fittings and a quality leak valve.

Regards,
Woody

Woody White
BWXT Services:
http://www.bwxt.com/bwxt.html
My Site:
http://woody.white.home.att.net

-----Original Message-----
} From: Ritchie Sims [mailto:r.sims-at-auckland.ac.nz]
Sent: Wednesday, May 12, 2004 1:45 AM
To: Microscopy-at-sparc5.microscopy.com


I'm sorry that this is a bit of a deviation, but, knowing how diverse this
list is, somebody
may be able to help:-

I have a Stanton-Redcroft Thermal Analyser, an STA-1600, the company in its
last
years was acquired by Polymer Laboratories and then by Rheometrics, and I'm
interested in exploring the possibility of hooking analytical instruments,
maybe NDIR or
even mass-spec, to the gas outlet line, to analyse the evolved gases (what
was referred
to some time ago, so nicely, by Stanton-Redcroft as "hyphenated
techniques").

Does anyone have such a setup?

Presumably one would use a heated PTFE interface tube, but the furnace
internal
volume is fairly large, and not necessarily efficiently swept by the 50
ml/min or so of
purge gas, and what I'm wondering about is the time resolution of such EGA
in this
instrument.

cheers

rtch

--
Ritchie Sims Ph D Phone : 64 9 3737599 ext
87713
Microanalyst Fax : 64 9 3737435
Department of Geology email :
r.sims-at-auckland.ac.nz
The University of Auckland
Private Bag 92019
Auckland
New Zealand



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 07:42:46 2004



From: qualityimages :      qualityimages-at-netrax.net
Date: Wed, 12 May 2004 08:48:26 -0400
Subject: [Microscopy] Re: RE: Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

shAf,
I've thought that for a long time but have yet to find a software-based
calibration that can use the 2 Cu peaks. Maybe the vendors can speak to
this.

As to Gary's question of why one needs standards, I have a hunch
they're not looking for EDS calibration standards (a piece of Cu tape on
an Al stub works just fine) but quantitative analysis standards for high
quality quant work. Despite the software advances in standardless
analysis, everyone I've talked to who really knows x-ray says that
standardless quant just doesn't have the accuracy or precision of
standards-based analysis.

Ken Converse
owner
Quality Images
third party SEM service
Delta, PA

michael shaffer wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Wed May 12 08:48:21 2004



From: sp-bf-at-physik.upb.de (by way of MicroscopyListserver)
Date: Wed, 12 May 2004 08:54:46 -0500
Subject: [Microscopy] viaWWW: repair of Hitachi H-600

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (sp-bf-at-physik.upb.de) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, May 12, 2004 at 05:24:26
---------------------------------------------------------------------------

Email: sp-bf-at-physik.upb.de
Name: Bettina Friedel

Organization: University of Paderborn

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi all!

Our physics group got a Hitachi H-600 a few years ago from biology group. We repaired it and there was no problem. But now the wire from the linkage of the specimen chamber has broken. It jumped out from the guide rolls and now we cannot move the specimen in one direction. "Nissei Sanyo Europe",the Hitachi Service Center here, demands 3500Ä for repair of this silly wire.
I hope anyone can help me!!! Is there a plan (blueprint?) how to thread the wire through the rolls, so that we could repair it ourself, not perfect but in this way, that moving of the specimen would be possible?

Bettina Friedel

---------------------------------------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 09:33:30 2004



From: Michal Jarnik :      M_Jarnik-at-fccc.edu
Date: Wed, 12 May 2004 10:39:45 -0400
Subject: [Microscopy] Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am trying to label the GFP in yeast on either Lowicryl sections or Tokuyashu cryosections. So far, we have had rather poor labeling/high background. I would appreciate any recommendations on a particular commercially available antibody and the fixation protocol.


Also, we would need to label a biotin-tagged protein. We were trying Streptavidin - gold, but to little success. Is there a good anti-biotin antibody (or, even better, anti-biotin/gold conjugate) people have used
for this purpose before?

Any recommendations would be highly appreciated.

Michael Jarnik,
EM Facility
Fox Chase Cancer Center
Philadelphia, PA




From MicroscopyL-request-at-ns.microscopy.com Wed May 12 10:50:57 2004



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Wed, 12 May 2004 10:57:06 -0500
Subject: [Microscopy] Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Just guessing, but I would suppose the energy calibration is a rather
simplistic algorithm. I am pretty sure that is much easier to work with a
symmetric K peak and pick off the centroid than it is to deconvolute the L
peaks to find the energy of the alpha line.

Presumably the effect of the background slope and curvature is negligible
compared to the peak intensity if all they have to do is get a peak energy.

Warren

At 04:57 AM 5/12/2004, "michael shaffer" wrote:
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 12:36:06 2004



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Wed, 12 May 2004 12:41:34 -0500
Subject: [Microscopy] Re: Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

michael

i', sorry, but your second question is not clear to me. are you simply
trying to label your protein with gold, or is this a detection step
where you have already used the antibody to react with its antigen and
now want to see if where the antigen is? also, do you have any antibody
which is not tagged with biotin and that you could use for tagging with
either nanogold or directly onto larger gold labels? finally, could you
do an indirect reaction using gold tagged antibody to the host species
antibody you are using, eg, say the biotin-antibody is an IgG species,
raised in mouse, could you come back with colloidal gold:anti mouse IgG?

paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-954
Cell:204-781-1502
Fax:204-789-392



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 12:38:32 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Wed, 12 May 2004 10:49:29 -0700
Subject: [Microscopy] Re: RE: Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 12, 2004, at 2:57 AM, michael shaffer wrote:

} Out of curiosty, why would a calibration use these
} 2 peaks rather than CuK and CuL?? Understandably,
} the L-line would be a convolution of the L-family,
} but I would think the Al K line is in a more difficult
} location relative to the slope in the continuum(???)
} Is there any preference with respect to EDX window
} type? (e.g., Be v UTW)
}
} I've thought that for a long time but have yet to find a
} software-based calibration that can use the 2 Cu peaks. Maybe the
} vendors can speak to this.

Dear Mike and Ken,
Since the Al k-line is at 1.4 keV and the Cu l-line(s) is at ~1 keV,
the Al is more readily separated from the continuum. I don't know of a
difference between Be and UTW calibrations, since it is the response of
the crystal that is being measured, but, perhaps there is an advantage
to using a lower-energy line with the UTW.
The TN2000 had software that could use any two lines. One simply
entered the energies and ran the calibration routine. I guess someone
must have "improved" the system by entering the line energies for us.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Wed May 12 12:54:20 2004



From: Michal Jarnik :      M_Jarnik-at-fccc.edu
Date: Wed, 12 May 2004 13:59:54 -0400
Subject: [Microscopy] Re: Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sorry, I would try to make it clearer. We need to localize a viral
protein which has been biotin-tagged and expressed in cell culture. I
expected it should be no problem to label it with Streptavidin
conjugate, but it is not that easy, it seems (even according to some
other people). So, as we do not have a good antibody to the viral
protein, I am considering an indirect labeling with anti-biotin
antibody and a secondary conjugate. There are tons of commercially
available anti-biotin antibodies, but certainly not all of them will
work with EM protocols - so I was interested whether somebody has
experience with some of these.

Thanks,

Michael

paul r hazelton wrote:

} michael
}
} i', sorry, but your second question is not clear to me. are you simply
} trying to label your protein with gold, or is this a detection step
} where you have already used the antibody to react with its antigen and
} now want to see if where the antigen is? also, do you have any antibody
} which is not tagged with biotin and that you could use for tagging with
} either nanogold or directly onto larger gold labels? finally, could you
} do an indirect reaction using gold tagged antibody to the host species
} antibody you are using, eg, say the biotin-antibody is an IgG species,
} raised in mouse, could you come back with colloidal gold:anti mouse IgG?
}
} paul
}
} Paul R. Hazelton, PhD
} Electron Microscope Unit
} University of Manitoba
} Department of Medical Microbiology
} 531 Basic Medical Sciences Building
} 730 William Avenue
} Winnipeg, Manitoba, Canada, R3E 0W3
} e-mail: paul_hazelton-at-umanitoba.ca
} Phone:204-789-3313
} Pager:204-931-954
} Cell:204-781-1502
} Fax:204-789-392
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 13:22:55 2004



From: rpowell-at-nanoprobes.com (by way of MicroscopyListserver)
Date: Wed, 12 May 2004 13:29:23 -0500
Subject: [Microscopy] viaWWW: Anti-GFP and anti-biotin Abs for TEM (vendor reply)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (rpowell-at-nanoprobes.com) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, May 12, 2004 at 12:22:51
---------------------------------------------------------------------------

Email: rpowell-at-nanoprobes.com
Name: Rick Powell

Organization: Nanoprobes, Incorporated

Title-Subject: [Microscopy] Anti-GFP and anti-biotin Abs for TEM (vendor reply)

Question: Hello Michael:

[Vendor reply...] We (Nanoprobes, Inc,) make an anti-biotin-IgG gold conjugate using our 1.4 nm Nanogold particle; we also offer streptavidin-Nanogold. You don't say why the streptavidin-gold didn't work - if you could let us have some details, we may have some suggestions.

Hope this is useful,

Rick Powell
Nanoprobes, Inc.



---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed May 12 13:24:42 2004



From: Tomic, Peter \(Peter\) :      ptomic-at-agere.com
Date: Wed, 12 May 2004 14:30:04 -0400
Subject: [Microscopy] Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

In a discussion I had with a service rep for EDAX Inc., the calibration is nothing more than setting the A/D [Analog to digital] converter's scale and offset. It doesn't change anything with respect to the detector itself.

-----Original Message-----
} From: Warren E Straszheim [mailto:wesaia-at-iastate.edu]
Sent: Wednesday, May 12, 2004 11:57 AM
To: Microscopy-at-msa.microscopy.com

Just guessing, but I would suppose the energy calibration is a rather
simplistic algorithm. I am pretty sure that is much easier to work with a
symmetric K peak and pick off the centroid than it is to deconvolute the L
peaks to find the energy of the alpha line.

Presumably the effect of the background slope and curvature is negligible
compared to the peak intensity if all they have to do is get a peak energy.

Warren

At 04:57 AM 5/12/2004, "michael shaffer" wrote:
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From MicroscopyL-request-at-ns.microscopy.com Wed May 12 13:26:36 2004



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Wed, 12 May 2004 14:32:15 -0400
Subject: [Microscopy] Re: Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Michael,

I've had great results on cryo-sections with a rabbit anti-GFP antibody
from
Molecular Probes: cat# 11122
Works well at dilutions ranging from 1:50-1:200.
Low background.
Hope it works for you.
Best

Marc


On Wednesday, May 12, 2004, at 10:39 AM, Michal Jarnik wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} I am trying to label the GFP in yeast on either Lowicryl sections or
} Tokuyashu cryosections. So far, we have had rather poor labeling/high
} background. I would appreciate any recommendations on a particular
} commercially available antibody and the fixation protocol.
}
} Also, we would need to label a biotin-tagged protein. We were trying
} Streptavidin - gold, but to little success. Is there a good
} anti-biotin antibody (or, even better, anti-biotin/gold conjugate)
} people have used for this purpose before?
} Any recommendations would be highly appreciated.
} Michael Jarnik,
} EM Facility
} Fox Chase Cancer Center
} Philadelphia, PA
}
}
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 13:32:23 2004



From: Sharon_Goresh-at-engelhard.com
Date: Wed, 12 May 2004 14:19:43 -0400
Subject: [Microscopy] LIMS for the Microscopy Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html





Fellow Microscopists,

The Microscopy Lab in our Research Center has 3 SEM's, 1 TEM and 1 EMPA
(with a second due in November) in addition to 4 XRD instruments. Our
main function has been as a problem solving facility to our researchers as
well as our field service engineers. Data produced by our lab is usually
issued as reports with images, spectra and graphs as attachments.
Management has directed us to implement a LIMS by the end of the year.
Their requirements include the ability of the researchers and field service
engineers to log samples in from their locations and the tracking of the
sample from the arrival in our lab to the issuance of data. It must also
be able to archive the data as well. For us this means literally hundreds
of images, spectra and reports each month that must be categorized and
saved. The researchers and engineers must also be able to access the data
from their location. This will probably mean a web based system since our
facilities cover the globe.


Has anyone out there been faced with a similar request? Our main hurdle is
convincing management that ours is not a "sample in - number out" type of
lab, and that most LIMS packages out there are designed for that type of
simple operation. If you have implemented such a system, was it an off the
shelf package, a custom designed package or a combination? If you want to
respond to me directly that would be fine. I am not interested in "system
bashing" just some helpful suggestions on how to make management happy
without losing my sanity.

Sharon Goresh
Chemist
sharon.goresh-at-engelhard.com



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 15:04:11 2004



From: Tom Phillips :      phillipst-at-missouri.edu
Date: Wed, 12 May 2004 15:11:47 -0500
Subject: [Microscopy] Re: Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

we like the bethyl labs mouse anti-biotin then follow that with goat
anti-mouse conjugated to gold. it is much better than streptavidin
gold. i am looking for a gfp ab so let me know what works.

At 10:39 AM 5/12/2004 -0400, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thomas E. Phillips, PhD
Professor of Biological Sciences
Director, Molecular Cytology Core
3 Tucker Hall
University of Missouri
Columbia, MO 65211-7400

573-882-4712 (office)
573-882-0123 (fax)
PhillipsT-at-missouri.edu




From MicroscopyL-request-at-ns.microscopy.com Wed May 12 15:26:13 2004



From: Paul Smith :      psmith-at-rontecusa.com
Date: Wed, 12 May 2004 17:52:41 -0400
Subject: [Microscopy] Re: RE: Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


The Moran Scientific EDS software can use the Cu L and K peaks (or any others the
user chooses) for the peak position calibration. I do this.

It would be a bit more correct not to use L peaks, because, as Shaf said, it's an
unresolved group of lines, so the resolution that gets calculated at the low energy end
isn't quite correct, but unless you're using virtual standards, that doesn't matter.

But hey, even the Al K peak is the Ka1 plus the Ka2 plus the Kb anyway, on an EDS.

cheers

rtch



Date sent: Wed, 12 May 2004 08:48:26 -0400
} From: qualityimages {qualityimages-at-netrax.net}
To: michael shaffer {michael-at-shaffer.net} ,
Microscopy {Microscopy-at-Sparc5.Microscopy.Com}

Dear Michael,
I just checked, and my Quartz XOne system can use any two peaks, including Cu Ka
and Cu La, to calibrate the system. The tradition may be to use Al Ka and Cu La
because those two elements are easy to find in most EM labs and the old
Be-window detectors did not always see Cu La well. Of course, the accuracy of
your calibration will depend on the separation of your peaks and their accurate
location. Ka peaks are sharper than La peaks, so it may improve your calibration
accuracy to use two Ka peaks.
Disclaimer: I act as a EDS consultant to Quartz Imaging Corp. on occasion.
Regards,
Mary Mager
Electron Microscopist
Department of Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca

----- Original Message -----
} From: "michael shaffer" {michael-at-shaffer.net}
To: "Gary Gaugler" {gary-at-gaugler.com}
Cc: "MSA listserver" {Microscopy-at-MSA.Microscopy.Com}
Sent: Wednesday, May 12, 2004 2:57 AM

Michael,

In response to your question, most, if not all, EDX manufacturers today
including RÖNTEC find it necessary to use only one specified energy line
plus the zero peak. This is because linearity is so high, that a 3-point
calibration (two specified spectral lines + zero peak) for almost all cases
is not beneficial.

Martin Rohde
RÖNTEC GmbH
RÖNTEC USA, Inc.
www.rontec.com

} On May 12, 2004, at 2:57 AM, michael shaffer wrote:

} Out of curiosty, why would a calibration use these
} 2 peaks rather than CuK and CuL?? Understandably,
} the L-line would be a convolution of the L-family,
} but I would think the Al K line is in a more difficult
} location relative to the slope in the continuum(???)
} Is there any preference with respect to EDX window
} type? (e.g., Be v UTW)
} }
} I've thought that for a long time but have yet to find a
} software-based calibration that can use the 2 Cu peaks. Maybe the
} vendors can speak to this.





From MicroscopyL-request-at-ns.microscopy.com Wed May 12 17:04:32 2004



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Wed, 12 May 2004 18:09:30 -0400
Subject: [Microscopy] Re: Re: Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Michael,

It's me again.
Actually I have used an anti-biotin that
worked well for immuno-EM.
It's from Rockland, cat # 200-4198.
It's a rabbit antibody, IgG fraction.

I too had problems using streptavidin-gold.
The reason for this, I think, is a lack of stability of
the conjugate. Don't forget that when you purchase
a gold conjugate, it might have sat for months on
a refrigerator shelve and the proteins have either
been degraded or fallen off the gold particles!

In my experience, good sources for antibodies
that work at the EM level are Cappel (bought a
a few years back by ICN - now called MP
Biomedicals), Rockland (to a lesser extent),
and Molecular Probes. I only use protein A
gold for my labelings, and I certainly avoid
gold conjugates from Ted Pella that have not
worked for me even once in the last 5 years!
Good luck

Marc


On Wednesday, May 12, 2004, at 01:59 PM, Michal Jarnik wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Sorry, I would try to make it clearer. We need to localize a viral
} protein which has been biotin-tagged and expressed in cell culture. I
} expected it should be no problem to label it with Streptavidin
} conjugate, but it is not that easy, it seems (even according to some
} other people). So, as we do not have a good antibody to the viral
} protein, I am considering an indirect labeling with anti-biotin
} antibody and a secondary conjugate. There are tons of commercially
} available anti-biotin antibodies, but certainly not all of them will
} work with EM protocols - so I was interested whether somebody has
} experience with some of these.
}
} Thanks,
}
} Michael
}
} paul r hazelton wrote:
}
} } michael
} }
} } i', sorry, but your second question is not clear to me. are you
} } simply
} } trying to label your protein with gold, or is this a detection step
} } where you have already used the antibody to react with its antigen and
} } now want to see if where the antigen is? also, do you have any
} } antibody
} } which is not tagged with biotin and that you could use for tagging
} } with
} } either nanogold or directly onto larger gold labels? finally, could
} } you
} } do an indirect reaction using gold tagged antibody to the host species
} } antibody you are using, eg, say the biotin-antibody is an IgG species,
} } raised in mouse, could you come back with colloidal gold:anti mouse
} } IgG?
} }
} } paul
} }
} } Paul R. Hazelton, PhD
} } Electron Microscope Unit
} } University of Manitoba
} } Department of Medical Microbiology
} } 531 Basic Medical Sciences Building
} } 730 William Avenue
} } Winnipeg, Manitoba, Canada, R3E 0W3
} } e-mail: paul_hazelton-at-umanitoba.ca
} } Phone:204-789-3313
} } Pager:204-931-954
} } Cell:204-781-1502
} } Fax:204-789-392
} }
} }
}
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 17:44:13 2004



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Thu, 13 May 2004 12:33:02 +1200
Subject: [Microscopy] Pinning down the drift

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I believe the real object of this is to find two peaks separated far enough in energy to allow the A/D converter to calibrate over a broad range. The broader the less the error. From an electronic point of view, it doesn't matter which peaks are used, just that they have significant separation. The electronics doesn't know the difference. It goes from peak height to a position in energy which corresponds to a voltage which gates a clock. I'm babbling, only electrical engineers worry about these things.

Any EDX vendors care to comment?

Peter

-----Original Message-----
} From: Mary Mager [mailto:mager-at-interchange.ubc.ca]
Sent: Wednesday, May 12, 2004 4:31 PM
To: michael shaffer
Cc: Microscopy

Dear Michael,
I just checked, and my Quartz XOne system can use any two peaks, including Cu Ka
and Cu La, to calibrate the system. The tradition may be to use Al Ka and Cu La
because those two elements are easy to find in most EM labs and the old
Be-window detectors did not always see Cu La well. Of course, the accuracy of
your calibration will depend on the separation of your peaks and their accurate
location. Ka peaks are sharper than La peaks, so it may improve your calibration
accuracy to use two Ka peaks.
Disclaimer: I act as a EDS consultant to Quartz Imaging Corp. on occasion.
Regards,
Mary Mager
Electron Microscopist
Department of Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca

----- Original Message -----
} From: "michael shaffer" {michael-at-shaffer.net}
To: "Gary Gaugler" {gary-at-gaugler.com}
Cc: "MSA listserver" {Microscopy-at-MSA.Microscopy.Com}
Sent: Wednesday, May 12, 2004 2:57 AM

Dear Peter,
Your age is showing, the electronics are all purely digital, now. No A/D
converter needed. The only reason to use two Ka peaks instead of a Ka and an La
peak are that the Ka peaks are sharper and so the peak centroid is more
precisely located. The software calibration sets the zero point and the
amplifier gain to a fraction of a channel. With the newer, higher-resolution
detectors, more quant precision and lower detection limit is possible, but the
calibration is more critical, since a tiny shift of the peak throws off all the
quant calculations. I find my self calibrating every month where I used to do it
once a year with the old A/D system, but my results are much better.
Disclaimer: I act as a EDS consultant to Quartz Imaging Corp. on occasion.
Regards,
Mary Mager
Electron Microscopist
Department of Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca

----- Original Message -----
} From: "Tomic, Peter (Peter)" {ptomic-at-agere.com}
To: "Mary Mager" {mager-at-interchange.ubc.ca} ; "michael shaffer"
{michael-at-shaffer.net}
Cc: "Microscopy" {microscopy-at-MSA.microscopy.com}
Sent: Wednesday, May 12, 2004 3:49 PM

Dear Michael,
I just checked, and my Quartz XOne system can use any two peaks, including Cu Ka
and Cu La, to calibrate the system. The tradition may be to use Al Ka and Cu La
because those two elements are easy to find in most EM labs and the old
Be-window detectors did not always see Cu La well. Of course, the accuracy of
your calibration will depend on the separation of your peaks and their accurate
location. Ka peaks are sharper than La peaks, so it may improve your calibration
accuracy to use two Ka peaks.
Disclaimer: I act as a EDS consultant to Quartz Imaging Corp. on occasion.
Regards,
Mary Mager
Electron Microscopist
Department of Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca

----- Original Message -----
} From: "michael shaffer" {michael-at-shaffer.net}
To: "Gary Gaugler" {gary-at-gaugler.com}
Cc: "MSA listserver" {Microscopy-at-MSA.Microscopy.Com}
Sent: Wednesday, May 12, 2004 2:57 AM


} I find my self calibrating every month where I used to do it once a
} year with the old A/D system, but my results are much better.


Once a month!

Once a year!

I do it once or twice a day.

But that's Virgoans for you. We worry a lot.

cheers

rtch

--
Ritchie Sims Ph D Phone : 64 9 3737599 ext 87713
Microanalyst Fax : 64 9 3737435
Department of Geology email : r.sims-at-auckland.ac.nz
The University of Auckland
Private Bag 92019
Auckland
New Zealand



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 20:07:12 2004



From: John Brealey :      john.brealey-at-imvs.sa.gov.au
Date: Wed, 12 May 2004 22:43:32 +0930
Subject: [Microscopy] H-600 problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Bettina,
I have the user's manual for the Hitachi H-600.
I will email you the diagrams.
I remember my boss trying to fix this problem and it almost drove him crazy.

Good luck!

John Brealey
Medical Scientist
Electron Microscopy Unit
The Queen Elizabeth Hospital
IMVS - TQEH Pathology
Woodville, 5011
South Australia
(08) 82226612

Bettina Friedel wrote:

Email: sp-bf-at-physik.upb.de
Name: Bettina Friedel

Organization: University of Paderborn

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi all!

Our physics group got a Hitachi H-600 a few years ago from biology group. We
repaired it and there was no problem. But now the wire from the linkage of
the specimen chamber has broken. It jumped out from the guide rolls and now
we cannot move the specimen in one direction. "Nissei Sanyo Europe",the
Hitachi Service Center here, demands 3500Ä for repair of this silly wire.
I hope anyone can help me!!! Is there a plan (blueprint?) how to thread the
wire through the rolls, so that we could repair it ourself, not perfect but
in this way, that moving of the specimen would be possible?

Bettina Friedel

____________________________



From MicroscopyL-request-at-ns.microscopy.com Wed May 12 22:08:20 2004



From: Gary Gaugler :      gary-at-gaugler.com
Date: Wed, 12 May 2004 20:13:57 -0700
Subject: [Microscopy] Re: RE: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


That is nice, but I would like more justification about
why only one point is adequate. I do not buy into this
hypothesis.

EDAX uses Al and Cu peaks. I think that this works very well.
And, it uncovers subtle problems that otherwise would go
un-noticed. I cannot see how a one point calibration
is a calibration at all. Please explain this to me.

The two disparate Al-Cu peaks makes a lot of sense for
calibrating an EDS. In addition to this, I do C, O, F
and Al measurements with one stub. This is a separate
"standard." When the results shift, if they do, I am
concerned. Lately, they have shifted. And I am concerned.
EDAX is on the job. Cool.

The point is to establish a set of standards and
use them frequently. When the results change, contact
the manufacturer. They will not know your situation
unless you tell them.

gary g.


At 02:52 PM 5/12/2004, you wrote:
} Michael,
}
} In response to your question, most, if not all, EDX manufacturers today
} including RÖNTEC find it necessary to use only one specified energy line
} plus the zero peak. This is because linearity is so high, that a 3-point
} calibration (two specified spectral lines + zero peak) for almost all cases
} is not beneficial.
}
} Martin Rohde
} RÖNTEC GmbH
} RÖNTEC USA, Inc.
} www.rontec.com
}
} } On May 12, 2004, at 2:57 AM, michael shaffer wrote:
}
} } Out of curiosty, why would a calibration use these
} } 2 peaks rather than CuK and CuL?? Understandably,
} } the L-line would be a convolution of the L-family,
} } but I would think the Al K line is in a more difficult
} } location relative to the slope in the continuum(???)
} } Is there any preference with respect to EDX window
} } type? (e.g., Be v UTW)
} } }
} } I've thought that for a long time but have yet to find a
} } software-based calibration that can use the 2 Cu peaks. Maybe the
} } vendors can speak to this.




From MicroscopyL-request-at-ns.microscopy.com Thu May 13 02:22:32 2004



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Thu, 13 May 2004 09:26:13 +0200
Subject: [Microscopy] RE: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Intersting !

The topic shifted a bit but with interesting points !

About that calibration questions, my few (Euro) cents :

No ADC ? Of coarse, an ADC is still necessary, but just after the
(still analogic) preamp, and no more after the hole ( former analogic)
counting unit.

Two points or more for energy calibration ? First point, the zero.
I suppose that all today sold systems, like that we have, have a dynamic
zero measurement, done on the noise peak. So nothing to do with the zero,
it is continously corrected. I suppose too that the energie range of the
(numerical) MCA is linear. So one (measured) peak in the midlle of the
range, that is between 6 and 12 keV will be good, as said a k line will be
sharper. Fe, Co or Cu... On our first Tracor NS 880, it was the first job
every morning, to play with the the gain and offset trimpot, on two lines.
With 150 eV resolution and a Be window, Al K was better tha Cu L (+ Cu K
for high energy). A good way to put your nerves in condition to welcome
your boss with a smile, when he came with a stupid question !

How often is it necessary to calibrate it ? Depends of the
stability of the system and of the accuracy wanted. But realy, the need of
a new calibration will be easily seen when one look the currently acquired
spectra. Particulary when one work with the same family of coumpound. A
shift in energy, whith a good statistic, if not been quantified, can be
well detected with the eyes ! It's too with the eyes on the spectrum, that
one can apreciated the real presence of a trace element, even when the
soft say that you have 0.000005 % of element Xy (without overlapping, of
coarse) !


And the embeding of my standards ? Nothing more to suggest ? My tests with
the epoxy I have show that it is not fluid enough. So, much bulbes...

Jacques



J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr





From MicroscopyL-request-at-ns.microscopy.com Thu May 13 02:36:03 2004



From: Allen Sampson :      ars-at-sem.com
Date: Thu, 13 May 2004 02:42:04 -0500
Subject: [Microscopy] RE: Re: Re: RE: Re: EDS standards embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Resisted the urge to jump in before now, but can't any longer.

Your reasons for use of two K peaks is good, might I also add that K peaks
will also offer better statistics as their count rates will be higher, thus
distinguishing them better from noise and background. Most software these
days allows the use of any two peaks. The choice of calibration peaks
should be made on the materials you normally analyze. There can be
non-linearities in the detector, pre-amplifier and amplifier responses that
could be a problem if you choose peaks that are too widely separated.
Aluminum and copper are commonly used because they fall within the range
of elements commonly analyzed. If, on the other hand, you commonly are
analyzing lanthanide's or actinides, you might want to choose peaks of
higher atomic number.

Problem is that we still have to render the avalanche pulse output from a
silicon detector, that merges millions of electron transfers into a single
pulse who's voltage peak represents energy of the x-ray photon causing it,
to a digital format. That pretty much is the definition of an analog to
digital converter. This function is not abandoned in current designs, only
performed earlier in the electronics. There is still an A/D converter in
the signal chain, it is just faster and controlled by a DSP.

Modern designs, because of increased time resolution of processors and A/D
converters, use digital signal processors (DSP) to provide many of the
functions that were previously done with dedicated electronics. This
trade-off has its own set of problems. Firmware is now responsible for the
identification of peaks, dead time and pulse pile-up and there may be cases
where that firmware could use a little tuning. This trend has resulted in
manufacturer's claims of greater counting rates and resolution but also
results in greater uncertainties as manufacturers strive to meet those
claims. If you are finding that you have to calibrate your EDS system more
frequently that is probably because of those uncertainties.

The recent trend has been to blame the inadequacies of EDS analyses on the
interpretation of the signals from the detectors rather than those signals
themself. In truth, we seem to have, for the moment, stretched the limit
of detector technology and perhaps over-stretched the detection of those
signals. The need for more frequent calibration may well be because the
state of the art in pulse measurement has exceeded the state of the art in
detector technology.

This thread has bifurcated into one that addresses the basic energy
calibration of the detector and electronics and one that addresses the
similar matrix calibrations required for good quantitative analysis. In
regards to that, let me say that these are two very different, yet
interrelated, areas. These are some of the most accurate measurements made
in common laboratory settings and deserve some common understanding. In
one case, we are calibrating our instrumentation to simplistic physical
properties we understand. In the other, we are accommodating our
instruments to complex interactions that we can't completely explain.

EDS calibration involves finding two reliable points that we can use to
establish a linear correction to the corresponding digital values that are
delivered to the computer doing the analysis. While the underlying
assumption is that this is a linear relationship, that's not necessarily
so. That assumption is generally made in the software that then uses this
data to produce results that claim to be quantitative using either
standards based or standardless algorithms.

Similar matrix calibration involves the use of well characterized standards
that are of a range of composition similar to that expected in the unknown
sample to provide a refinement of the simplistic linear corrections
mentioned above. These standards and corrections allow us to improve our
quantitative estimations by narrowing the range of linear corrections used.
In reality, the interelement interactions in any particular sample can not
be completely derived from the actions of their individual components at
the present time.

This is still an empirical science, like it or not. As such, we have to
continually accommodate our instrumentation to the task rather than to an
ideal. Those accommodations will continue to be a problem and a limitation
to the results we derive.



Allen R. Sampson
Advanced Research Systems
317 North 4th. Street
St. Charles, Illinois 60174

phone (630) 513-7093 fax (630) 513-7092 http://www.sem.com


On Wednesday, May 12, 2004 6:09 PM, Mary Mager
[SMTP:mager-at-interchange.ubc.ca] wrote:
}
}
}
------------------------------------------------------------------------
------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
-------
}
} Dear Peter,
} Your age is showing, the electronics are all purely digital, now. No A/D
} converter needed. The only reason to use two Ka peaks instead of a Ka and
an La
} peak are that the Ka peaks are sharper and so the peak centroid is more
} precisely located. The software calibration sets the zero point and the
} amplifier gain to a fraction of a channel. With the newer,
higher-resolution
} detectors, more quant precision and lower detection limit is possible,
but the
} calibration is more critical, since a tiny shift of the peak throws off
all the
} quant calculations. I find my self calibrating every month where I used
to do it
} once a year with the old A/D system, but my results are much better.
} Disclaimer: I act as a EDS consultant to Quartz Imaging Corp. on
occasion.
} Regards,
} Mary Mager
} Electron Microscopist
} Department of Materials Engineering
} University of British Columbia
} 6350 Stores Road
} Vancouver, B.C. V6T 1Z4
} CANADA
} tel: 604-822-5648
} e-mail: mager-at-interchange.ubc.ca
}
} ----- Original Message -----
} } From: "Tomic, Peter (Peter)" {ptomic-at-agere.com}
} To: "Mary Mager" {mager-at-interchange.ubc.ca} ; "michael shaffer"
} {michael-at-shaffer.net}
} Cc: "Microscopy" {microscopy-at-MSA.microscopy.com}
} Sent: Wednesday, May 12, 2004 3:49 PM
} Subject: [Microscopy] RE: Re: RE: Re: EDS standards embedding
}
}
} I believe the real object of this is to find two peaks separated far
enough in
} energy to allow the A/D converter to calibrate over a broad range. The
broader
} the less the error. From an electronic point of view, it doesn't matter
which
} peaks are used, just that they have significant separation. The
electronics
} doesn't know the difference. It goes from peak height to a position in
energy
} which corresponds to a voltage which gates a clock. I'm babbling, only
} electrical engineers worry about these things.
}
} Any EDX vendors care to comment?
}
} Peter
}
} -----Original Message-----
} } From: Mary Mager [mailto:mager-at-interchange.ubc.ca]
} Sent: Wednesday, May 12, 2004 4:31 PM
} To: michael shaffer
} Cc: Microscopy
} Subject: [Microscopy] Re: RE: Re: EDS standards embedding
}
}
}
}
}
------------------------------------------------------------------------
------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
-------
}
} Dear Michael,
} I just checked, and my Quartz XOne system can use any two peaks,
including Cu Ka
} and Cu La, to calibrate the system. The tradition may be to use Al Ka and
Cu La
} because those two elements are easy to find in most EM labs and the old
} Be-window detectors did not always see Cu La well. Of course, the
accuracy of
} your calibration will depend on the separation of your peaks and their
accurate
} location. Ka peaks are sharper than La peaks, so it may improve your
calibration
} accuracy to use two Ka peaks.
} Disclaimer: I act as a EDS consultant to Quartz Imaging Corp. on
occasion.
} Regards,
} Mary Mager
} Electron Microscopist
} Department of Materials Engineering
} University of British Columbia
} 6350 Stores Road
} Vancouver, B.C. V6T 1Z4
} CANADA
} tel: 604-822-5648
} e-mail: mager-at-interchange.ubc.ca
}
} ----- Original Message -----
} } From: "michael shaffer" {michael-at-shaffer.net}
} To: "Gary Gaugler" {gary-at-gaugler.com}
} Cc: "MSA listserver" {Microscopy-at-MSA.Microscopy.Com}
} Sent: Wednesday, May 12, 2004 2:57 AM
} Subject: [Microscopy] RE: Re: EDS standards embedding
}
}
} }
} }
} }
------------------------------------------------------------------------
------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} } To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} } On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
------------------------------------------------------------------------
------
} -
} }
} } Gary writes ...
} }
} } } The "normal" calibration of EDS is done at Al
} } } K alpha (1.486KeV) and Cu K alpha (8.040KeV).
} } } ...
} }
} } Out of curiosty, why would a calibration use these
} } 2 peaks rather than CuK and CuL?? Understandably,
} } the L-line would be a convolution of the L-family,
} } but I would think the Al K line is in a more difficult
} } location relative to the slope in the continuum(???)
} } Is there any preference with respect to EDX window
} } type? (e.g., Be v UTW)
} }
} } cheerios ... shAf :o)
} } Avalon Peninsula, Newfoundland
} } www.micro-investigations.com (in progress)
} }
} }
}
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 02:47:11 2004



From: Peter Van Osta :      pvosta-at-maia-scientific.com
Date: Thu, 13 May 2004 09:53:08 +0200
Subject: [Microscopy] Dealing with image sequences

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,

Most of us at some moment have to deal with the capture and storage of
image sequences, either as part of a video-experiment or when doing a
video time-lapse. Until 3 years ago, I did my video-microscopy on a
Silicon Graphics based system. The nice features of SGI were that we
could do multiple videostreams in parallel, which allowed us to do
multiposition time-lapse experiments.

However, three years ago I changed to another Unix-platform (Linux) and
to my unpleasant surprise, it was not so trivial anymore to keep
multiple video-streams in the air during an experiment and there were
some other caveats too.

In the end I decided that the only thing I needed was a simple way to
concatenate individual frames into a filestream, as for me the most
important issue is to put the frames where I can find them afterwards
for analysis, not so much for looking a them over anetwork at high
speed. The format should be simple, no overhead and frame-oriented, so
even mixing frames of different types should be possible. Another thing
I wanted to have, was that it should be possble to store 3D-frame
movie-sequences. Lossy compression schemes are not so beneficial for
image anlsysis aftwerwards, as they manipulate the frame content which
causes artefacts in the analysis.

The drawback of this frame-oriented approach is that some lossy
compression techniques, which require multiple frame comparison, are not
possible.

I would like to know if there are other people, doing video microscopy
who are also interested in this approach, maybe we could create a simple
software-library (ANSI C) which we can link with software we use, so we
can have a video-system designed and optimized for video-microscopy
(analysis oriented instead of visualization-oriented) ?

Regards,

Peter Van Osta

Director Imaging
MAIA SCIENTIFIC (formerly Union Biometrica NV)
Cipalstraat 3
B-2440 Geel
Belgium
Tel.: +32-(0)14 570 620
Fax.: +32-(0)14 570 621
Email: pvosta-at-maia-scientific.com
Website: www.maia-scientific.com
A Harvard Bioscience Company



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 03:08:48 2004



From: Richard Beanland :      richard.beanland-at-bookham.com
Date: Thu, 13 May 2004 09:13:42 +0100
Subject: [Microscopy] LIMS for the Microscopy Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sharon,
it is possible to spend a lot of money on these systems. However, I
have yet to find one which offers significantly more than can be done with
MSExcel spreadsheets.

MSExcel spreadsheets on a server can be used in a 'shared' format, where
many people can open the same file without that irritating 'This file is
being used by...' error. In the system we have here, the files can only be
written to by a member of the analysis group, but read by anyone. People
have to come here to give us a sample anyway, so this is not a problem.
Each sample gets a line in the spreadsheet with the usual fields (date,
name, analysis needed etc.). When someone is working on a job, they put
their initials in the box for that analysis type; when the job is done, a
date goes in the box instead. The spreadsheet is updated every time someone
saves it. Each column has an 'Autofilter' at the top which means they can
be searched for any of the fields.
This 'master' spreadsheet contains hyperlinks to the electronic log books
for each microscope or technique (also shared MSExcel spreadsheets). These
are updated as the analysis happens, and here each image has a line giving
sample, analysis details, operator, customer, etc.. Each line in these
'technique' spreadsheets has a hyperlink to the data file. Again, each
column has an 'Autofilter' at the top which means they can be searched for
any of the fields.

As a user, it is easy to operate (that 'fill down' feature of Excel allows
you to copy the provious line and change one box without having to make 10
entries for each image). The only discipline needed is to keep saving the
files frequently (autosaving doesn't work, if two computers try to save at
the same time they can lock up). Our customers seem to be happy too - they
know the status of their job and can find the data without having to hassle
us. We rely on the standard software on each user's PC to be able to open
the images.

I'm not sure how this could be implemented on a web-based system, but for a
server it works very well. We keep track of about 4000 samples a year - and
all the data files they generate - with this system, and it usually only
takes a few minutes to find old data even from a vague description. I have
yet another spreadsheet which allows me to count how many jobs have been
done, are left to do, etc., for each analysis type. The nice thing about
spreadsheets is that they are totally 'customizable', so you only keep track
of what you think is important.

I'd be happy to send you an example of the files if you like. As a 'free'
solution it is hard to beat!

Cheers

Richard

_______________________________
Richard Beanland
Analytical Services
Bookham Technology plc
Caswell,
Towcester,
Northamptonshire NN12 8EQ
UK
Tel: +44 (0) 1327 356362
Fax: +44 (0) 1327 356775
http://www.bookham.com
All incoming mail is filtered for spam. If you do not receive a reply from
me, do not assume I have received your mail.


-----Original Message-----
} From: Sharon_Goresh-at-engelhard.com [mailto:Sharon_Goresh-at-engelhard.com]
Sent: 12 May 2004 19:20
To: Microscopy-at-MSA.Microscopy.Com





Fellow Microscopists,

The Microscopy Lab in our Research Center has 3 SEM's, 1 TEM and 1 EMPA
(with a second due in November) in addition to 4 XRD instruments. Our
main function has been as a problem solving facility to our researchers as
well as our field service engineers. Data produced by our lab is usually
issued as reports with images, spectra and graphs as attachments.
Management has directed us to implement a LIMS by the end of the year.
Their requirements include the ability of the researchers and field service
engineers to log samples in from their locations and the tracking of the
sample from the arrival in our lab to the issuance of data. It must also
be able to archive the data as well. For us this means literally hundreds
of images, spectra and reports each month that must be categorized and
saved. The researchers and engineers must also be able to access the data
from their location. This will probably mean a web based system since our
facilities cover the globe.


Has anyone out there been faced with a similar request? Our main hurdle is
convincing management that ours is not a "sample in - number out" type of
lab, and that most LIMS packages out there are designed for that type of
simple operation. If you have implemented such a system, was it an off the
shelf package, a custom designed package or a combination? If you want to
respond to me directly that would be fine. I am not interested in "system
bashing" just some helpful suggestions on how to make management happy
without losing my sanity.

Sharon Goresh
Chemist
sharon.goresh-at-engelhard.com





=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.
No part of this message can be considered a request for goods or
services.
=======================================================================
Any questions about Bookham's E-Mail service should be directed to
postmaster-at-bookham.com.



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 05:16:19 2004



From: Peter van de Plas :      p.vandeplas-at-aurion.nl
Date: Thu, 13 May 2004 12:29:24 +0200
Subject: [Microscopy] Re: Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Michael,

A two-step detection method is indeed a good option to visualize your
biotinylated viral protein. Compared to the 2-step method the advantage
of a 1-step immunodetection method, as you initially suggested, will be
improved resolution. Your labeling density decreases however. As you
know labeling density is inversely related to the gold particle size.

Improved resolution combined with high sensitivity can be obtained when
using gold conjugates prepared with Ultra-Small gold particles. For
your application we would recommend Goat anti Biotin Ultra-Small. See
e.g. Müller-Höcker et al., (1998), Histochem. and Cell Biol. 109(2),
119-125.

When the protein of interest is located on the outside of the virus, in
other words the biotin is accessible for the antibodies, another option
may be the use of a pre-embedding labeling procedure. The use of
Ultra-Small gold conjugates is a prerequisite to successfully combine
appropriate labeling densities with a decent morphology. See e.g.
Decossas et al, (2003) J. of Comp. Neurology 462(3), 302-314 for
details on this technique.

Hope this is of help
Kind regards,
Peter

------------------------------------------
Michael Jarnik wrote:

Sorry, I would try to make it clearer. We need to localize a viral
protein which has been biotin-tagged and expressed in cell culture. I
expected it should be no problem to label it with Streptavidin
conjugate, but it is not that easy, it seems (even according to some
other people). So, as we do not have a good antibody to the viral
protein, I am considering an indirect labeling with anti-biotin
antibody and a secondary conjugate. There are tons of commercially
available anti-biotin antibodies, but certainly not all of them will
work with EM protocols - so I was interested whether somebody has
experience with some of these.

Thanks,

Michael

-----------
Peter van de Plas
Aurion
Costerweg 5
6702 AA Wageningen
The Netherlands
Phone: +31 317 497676
Fax: +31 317 415955



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 06:55:04 2004



From: Peter van de Plas :      p.vandeplas-at-aurion.nl
Date: Thu, 13 May 2004 13:55:14 +0200
Subject: [Microscopy] Re: Anti-GFP and anti-biotin Abs for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Michael,

A two-step detection method is indeed a good option to visualize your
biotinylated viral protein. Compared to the 2-step method the advantage
of a 1-step immunodetection method, as you initially suggested, will be
improved resolution. Your labeling density decreases however. As you
know labeling density is inversely related to the gold particle size.

Improved resolution combined with high sensitivity can be obtained when
using gold conjugates prepared with Ultra-Small gold particles. For
your application we would recommend Goat anti Biotin Ultra-Small. See
e.g. Müller-Höcker et al., (1998), Histochem. and Cell Biol. 109(2),
119-125.

When the protein of interest is located on the outside of the virus, in
other words the biotin is accessible for the antibodies, another option
may be the use of a pre-embedding labeling procedure. The use of
Ultra-Small gold conjugates is a prerequisite to successfully combine
appropriate labeling densities with a decent morphology. See e.g.
Decossas et al, (2003) J. of Comp. Neurology 462(3), 302-314 for
details on this technique.

Hope this is of help
Kind regards,
Peter

------------------------------------------
Michael Jarnik wrote:

Sorry, I would try to make it clearer. We need to localize a viral
protein which has been biotin-tagged and expressed in cell culture. I
expected it should be no problem to label it with Streptavidin
conjugate, but it is not that easy, it seems (even according to some
other people). So, as we do not have a good antibody to the viral
protein, I am considering an indirect labeling with anti-biotin
antibody and a secondary conjugate. There are tons of commercially
available anti-biotin antibodies, but certainly not all of them will
work with EM protocols - so I was interested whether somebody has
experience with some of these.

Thanks,

Michael

-----------
Peter van de Plas
Aurion
Costerweg 5
6702 AA Wageningen
The Netherlands
Phone: +31 317 497676
Fax: +31 317 415955



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 09:29:14 2004



From: rosscac-at-aol.com (by way of MicroscopyListserver)
Date: Thu, 13 May 2004 09:42:23 -0500
Subject: [Microscopy] viaWWW: Print processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (rosscac-at-aol.com) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
Thursday, May 13, 2004 at 07:52:40
---------------------------------------------------------------------------

Email: rosscac-at-aol.com
Name: Connie Cummings

Title-Subject: [Microscopy] [Filtered] Print processors

Question: I am interested in purchasing a print processor and was
wondering if anyone out there is using or has used either a Fujimoto
Roller Transport print processor CP51 or a JOBO printlab 3504 print
processor. I was looking for some feedback on these units on there
durablility and functionalitiy or any other tidbits! You can respond
back via email to me at rosscac-at-aol.com or at
connie.a.cummings-at-gsk.com.
Thanks in advance.
Connie

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu May 13 10:01:17 2004



From: sghoshro-at-NMSU.Edu
Date: Thu, 13 May 2004 09:07:05 -0600 (MDT)
Subject: [Microscopy] GFP survival

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi fellow microscopists,

A post doc is trying to detect GFP being expressed in plant leaves. She is
fixing leaves in formaldehyde and doing a standard paraffin embedding. Her
question is if GFP going to survive xylene treatments for paraffin
infiltration. She plans to section the paraffin embedded blocks and look
for GFP using fluorescence microscopy.

If anyone has an answer please reply back to me.

Thanks in advance.

Soumitra

*************************************************************
Soumitra Ghoshroy
College Associate Professor, Biology
Director, Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
http://www.cs.nmsu.edu/~biology/Faculty%20&%20Staff/Ghoshroy/Ghoshroy.htm
http://emldata.nmsu.edu/eml/


From MicroscopyL-request-at-ns.microscopy.com Thu May 13 10:16:44 2004



From: sylvain maury :      sylvain.maury-at-thalesgroup.com
Date: Thu, 13 May 2004 17:22:40 +0200
Subject: [Microscopy] graphite sputtering or coating

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

hi all!

a simple question :

Does anyone have ever used graphite for sputtering or coating a SEM
sample ?

maybe a question from a dummy boy :-) , doesn't it?

thanx for answers...

Sylvain MAURY



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 10:35:04 2004



From: Beth Richardson :      beth-at-plantbio.uga.edu
Date: Thu, 13 May 2004 11:38:58 -0400
Subject: [Microscopy] slide coating

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all,
Lazy me recommended pre-coated SuperFrostPlus slides to Anne Dunn a
post-doc working on antlions but the gut goo doesn't always survive her
processing protocol (see her message below).

So the question is - What is the preferred slide coating method?
Gelatin vs poly-L-lysine...or something else?

advice and recipes would be greatly appreciated.
thanks,
Beth

} Hi Beth,
}
} I am finding that when I use the superfrost plus slides for
} my cryosections that often if there is material inside the
} gut that it will wash away as I am fixing and dehydrating
} my sections before hybridization (and I am not seeing any
} bacteria in the gut area after hybridization). I have no idea
} whether it is just something that is going to happen with
} a fluid-feeding insect, or if it is the superfrost slides not
} being sticky enough. I don't seem to be having a
} problem with the tissue itself sticking.
}
} I was looking up ways to coat slides and found there is
} the gelatin method and the poly-L-lysine method.
} Unfortunately I am not finding any info that tells me if one
} is better than the other. Do you have an opinion? I was
} wondering if it is a case of one coating being better for
} certain applications than another?
}
} I was going to go ahead and dissect out a gut or two, fix
} and hybridize them, trying to see if I can find any
} bacteria-like objects in the gut smooshate...to let me
} know if the loss of the material in the gut is an issue or
} not, but wanted to start thinking/plotting if I need to coat
} some slides.

**********************************************************************
Beth Richardson
EM Lab Coordinator
Plant Biology Department
University of Georgia
Athens, GA 30602-7271

Phone - (706) 542-1790 & FAX - (706) 542-1805

"Between the two evils,
I always pick the one I never tried before". Mae West (1893-1980)
*******************************************************************

"And it's only the giving that makes you what you are".
Wond'ring Aloud, Jethro Tull (Aqualung)

************************************************************************
***




From MicroscopyL-request-at-ns.microscopy.com Thu May 13 10:52:50 2004



From: Rick A. Harris :      raharris-at-ucdavis.edu
Date: Thu, 13 May 2004 08:58:55 -0700
Subject: [Microscopy] Leica (Reichert) Ultacut T

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I'm looking for a shield to exclude air movement around the stage of my
Ultracut. I remember years ago having one for my Reichert OM-U3 but I
don't believe I got one for my Ultracut a few years ago. I've checked a
couple of the vendors and found nothing. I could find nothing at the
Leica site either. Does anyone have a source for these or will I have
to make my own?

Rick A. Harris, exDirector
Microscopy and Imaging Facility
Section of Molecular and Cellular Biology
1241 Life Sciences Addition
University of California
Davis, CA
530 752 2914
http://microscopy.mcb.ucdavis.edu
raharris-at-ucdavis.edu




From MicroscopyL-request-at-ns.microscopy.com Thu May 13 11:41:29 2004



From: Mike Bode :      mb-at-soft-imaging.com
Date: Thu, 13 May 2004 10:45:59 -0600
Subject: [Microscopy] LIMS for the Microscopy Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Sharon,

Richard is of course right -- you can spend a fortune on LIMS systems and
for most users it may be overkill. On the other hand, the Excel sheets that
Richard mentions may be too limited.

A third option, somewhat in the middle, could be to use software that
supports or can read the images and other data from your software and keep
track of those in an archive or database. That way you can search for
keyfields, you deal with the same software for all applications, and you can
distribute the images easily. Our software analySIS includes such a
database. We also have a web frontend for this database so you can deal with
the global issues you mentioned.

Please take a look at our website or contact me directly if you wolud like
to get more information.

Mike


Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Richard Beanland [mailto:richard.beanland-at-bookham.com]
Sent: Thursday, May 13, 2004 02:14
To: 'Sharon_Goresh-at-engelhard.com'
Cc: 'Microscopy-at-MSA.Microscopy.Com'

Sharon,
it is possible to spend a lot of money on these systems. However, I
have yet to find one which offers significantly more than can be done with
MSExcel spreadsheets.

MSExcel spreadsheets on a server can be used in a 'shared' format, where
many people can open the same file without that irritating 'This file is
being used by...' error. In the system we have here, the files can only be
written to by a member of the analysis group, but read by anyone. People
have to come here to give us a sample anyway, so this is not a problem.
Each sample gets a line in the spreadsheet with the usual fields (date,
name, analysis needed etc.). When someone is working on a job, they put
their initials in the box for that analysis type; when the job is done, a
date goes in the box instead. The spreadsheet is updated every time someone
saves it. Each column has an 'Autofilter' at the top which means they can
be searched for any of the fields.
This 'master' spreadsheet contains hyperlinks to the electronic log books
for each microscope or technique (also shared MSExcel spreadsheets). These
are updated as the analysis happens, and here each image has a line giving
sample, analysis details, operator, customer, etc.. Each line in these
'technique' spreadsheets has a hyperlink to the data file. Again, each
column has an 'Autofilter' at the top which means they can be searched for
any of the fields.

As a user, it is easy to operate (that 'fill down' feature of Excel allows
you to copy the provious line and change one box without having to make 10
entries for each image). The only discipline needed is to keep saving the
files frequently (autosaving doesn't work, if two computers try to save at
the same time they can lock up). Our customers seem to be happy too - they
know the status of their job and can find the data without having to hassle
us. We rely on the standard software on each user's PC to be able to open
the images.

I'm not sure how this could be implemented on a web-based system, but for a
server it works very well. We keep track of about 4000 samples a year - and
all the data files they generate - with this system, and it usually only
takes a few minutes to find old data even from a vague description. I have
yet another spreadsheet which allows me to count how many jobs have been
done, are left to do, etc., for each analysis type. The nice thing about
spreadsheets is that they are totally 'customizable', so you only keep track
of what you think is important.

I'd be happy to send you an example of the files if you like. As a 'free'
solution it is hard to beat!

Cheers

Richard

_______________________________
Richard Beanland
Analytical Services
Bookham Technology plc
Caswell,
Towcester,
Northamptonshire NN12 8EQ
UK
Tel: +44 (0) 1327 356362
Fax: +44 (0) 1327 356775
http://www.bookham.com
All incoming mail is filtered for spam. If you do not receive a reply from
me, do not assume I have received your mail.


-----Original Message-----
} From: Sharon_Goresh-at-engelhard.com [mailto:Sharon_Goresh-at-engelhard.com]
Sent: 12 May 2004 19:20
To: Microscopy-at-MSA.Microscopy.Com





Fellow Microscopists,

The Microscopy Lab in our Research Center has 3 SEM's, 1 TEM and 1 EMPA
(with a second due in November) in addition to 4 XRD instruments. Our
main function has been as a problem solving facility to our researchers as
well as our field service engineers. Data produced by our lab is usually
issued as reports with images, spectra and graphs as attachments.
Management has directed us to implement a LIMS by the end of the year.
Their requirements include the ability of the researchers and field service
engineers to log samples in from their locations and the tracking of the
sample from the arrival in our lab to the issuance of data. It must also
be able to archive the data as well. For us this means literally hundreds
of images, spectra and reports each month that must be categorized and
saved. The researchers and engineers must also be able to access the data
from their location. This will probably mean a web based system since our
facilities cover the globe.


Has anyone out there been faced with a similar request? Our main hurdle is
convincing management that ours is not a "sample in - number out" type of
lab, and that most LIMS packages out there are designed for that type of
simple operation. If you have implemented such a system, was it an off the
shelf package, a custom designed package or a combination? If you want to
respond to me directly that would be fine. I am not interested in "system
bashing" just some helpful suggestions on how to make management happy
without losing my sanity.

Sharon Goresh
Chemist
sharon.goresh-at-engelhard.com





=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.
No part of this message can be considered a request for goods or
services.
=======================================================================
Any questions about Bookham's E-Mail service should be directed to
postmaster-at-bookham.com.



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 12:27:22 2004



From: Paula Sicurello :      patpxs-at-gwumc.edu
Date: Thu, 13 May 2004 13:31:43 -0400
Subject: [Microscopy] Re: slide coating

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Beth,

We use silane to coat our slides, it is the standard for the histology labs in the pathology dept.

Silane = 3-aminopropyl-triethoxy-silane

Protocol:

Silane 20 ml
Actenoe 2000 ml

Dip precleaned slides in Silane solution for 2 minutes
Drain and wash in distilled water 2 times
Drain and dry at 60 degrees C for 1 hour
Store away from light and moisture.


Our sildes last for many months (6+), we make huge batches at a time.

I hope this helps you out of your sticky situation!

Paula :-)


Paula Sicurello
George Washington Univ. Medical Center
Electron Microscope Lab
Washington, DC 20037
202-994-2930 phone
202-994-2518 fax

} } } Beth Richardson {beth-at-plantbio.uga.edu} 05/13/04 11:38AM } } }


------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hi all,
Lazy me recommended pre-coated SuperFrostPlus slides to Anne Dunn a
post-doc working on antlions but the gut goo doesn't always survive her
processing protocol (see her message below).

So the question is - What is the preferred slide coating method?
Gelatin vs poly-L-lysine...or something else?

advice and recipes would be greatly appreciated.
thanks,
Beth

} Hi Beth,
}
} I am finding that when I use the superfrost plus slides for
} my cryosections that often if there is material inside the
} gut that it will wash away as I am fixing and dehydrating
} my sections before hybridization (and I am not seeing any
} bacteria in the gut area after hybridization). I have no idea
} whether it is just something that is going to happen with
} a fluid-feeding insect, or if it is the superfrost slides not
} being sticky enough. I don't seem to be having a
} problem with the tissue itself sticking.
}
} I was looking up ways to coat slides and found there is
} the gelatin method and the poly-L-lysine method.
} Unfortunately I am not finding any info that tells me if one
} is better than the other. Do you have an opinion? I was
} wondering if it is a case of one coating being better for
} certain applications than another?
}
} I was going to go ahead and dissect out a gut or two, fix
} and hybridize them, trying to see if I can find any
} bacteria-like objects in the gut smooshate...to let me
} know if the loss of the material in the gut is an issue or
} not, but wanted to start thinking/plotting if I need to coat
} some slides.

**********************************************************************
Beth Richardson
EM Lab Coordinator
Plant Biology Department
University of Georgia
Athens, GA 30602-7271

Phone - (706) 542-1790 & FAX - (706) 542-1805

"Between the two evils,
I always pick the one I never tried before". Mae West (1893-1980)
*******************************************************************

"And it's only the giving that makes you what you are".
Wond'ring Aloud, Jethro Tull (Aqualung)

************************************************************************
***







From MicroscopyL-request-at-ns.microscopy.com Thu May 13 13:19:34 2004



From: Dusevich, Vladimir :      DusevichV-at-umkc.edu
Date: Thu, 13 May 2004 13:25:35 -0500
Subject: [Microscopy]

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am going to buy scanner for TEM negatives and considering two models:
Microtek ArtixScan 1800f and Epson Expression 1680 Pro. Any advise and
comments are very welcome.

Thank you in advance,

Vladimir

Vladimir M. Dusevich, Ph.D.
Electron Microscope Lab Manager
3127 School of Dentistry
650 E. 25th Street
Kansas City, MO 64108-2784

Phone: (816) 235-2072
Fax: (816) 235-5524
Web: http://www.umkc.edu/dentistry/microscopy




From MicroscopyL-request-at-ns.microscopy.com Thu May 13 13:39:06 2004



From: aruna weberg :      aweberg-at-siumed.edu
Date: Thu, 13 May 2004 13:32:40 -0500
Subject: [Microscopy] Polaron Sputter Coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello,

I am looking for a replacement glass spacer for a Polaron E5100 series
II 'Cool' Sputter Coater. The external diameter etched on metal collar
base is 165.1 m/m. The internal diameter of the metal collar is 5.75".
The height of the glass spacer is 3 and 3/16". I have been working with
EBS to locate a replacement collar but would like to know if there are
any other sources I could contact.

Thank you in advance for your help

aruna



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 15:30:40 2004



From: Caroline Schooley :      schooley-at-mcn.org
Date: Thu, 13 May 2004 13:42:32 -0700
Subject: [Microscopy] Re: Leica (Reichert) Ultacut T

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Sylvain,
I use graphite, evaporated in a high-vacuum evaporator, to make SEM samples
conductive when I want to do EDS analysis. I use gold/palladium for SEM imaging
without analysis, but carbon coating to minimize any interference of the coating
with the EDS analysis.
Regards,
Mary Mager
Electron Microscopist
Department of Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca
----- Original Message -----
} From: "sylvain maury" {sylvain.maury-at-thalesgroup.com}
To: "Microscopy community" {microscopy-at-msa.microscopy.com}
Sent: Thursday, May 13, 2004 8:22 AM

} I'm looking for a shield to exclude air movement around the stage of my
} Ultracut. I remember years ago having one for my Reichert OM-U3 but I
} don't believe I got one for my Ultracut a few years ago. I've checked a
} couple of the vendors and found nothing. I could find nothing at the
} Leica site either. Does anyone have a source for these or will I have
} to make my own?
}
} Rick A. Harris, exDirector
} Microscopy and Imaging Facility
} Section of Molecular and Cellular Biology
} 1241 Life Sciences Addition
} University of California
} Davis, CA
} 530 752 2914
} http://microscopy.mcb.ucdavis.edu
} raharris-at-ucdavis.edu

Rick -

Make a hole in the bottom of a plastic veggie bag from your
supermarket & rubber-band it around the binocular objective. Ugly,
but it works better than a rigid plastic box.

Caroline
--
Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/marinelab.html


From MicroscopyL-request-at-ns.microscopy.com Thu May 13 16:09:34 2004



From: Vetrano, John S :      john.vetrano-at-pnl.gov
Date: Thu, 13 May 2004 14:15:02 -0700
Subject: [Microscopy] Scanner info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Short and sweet: about a year ago we bought a Microtek ArtixScan 2500f (firewire); a little different from the 1800f but I don't know how much. Anyway, we like it. Great dynamic range and reasonable software (we scan into a G4 Mac running OSX).

cheers, John

********
John S. Vetrano
Sr. Research Scientist
Materials Structure and Performance Group
Pacific Northwest National Laboratory
MSIN P8-16
P.O. Box 999
Richland, WA 99352
Phone: (509)372-0724 Fax: (509)376-6308
Email: mailto:john.vetrano-at-pnl.gov


} ----------
} From: Dusevich, Vladimir
} Sent: Thursday, May 13, 2004 11:25 AM
} To: microscopy-at-msa.microscopy.com
} Subject: [Microscopy]
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} I am going to buy scanner for TEM negatives and considering two models:
} Microtek ArtixScan 1800f and Epson Expression 1680 Pro. Any advise and
} comments are very welcome.
}
} Thank you in advance,
}
} Vladimir
}
} Vladimir M. Dusevich, Ph.D.
} Electron Microscope Lab Manager
} 3127 School of Dentistry
} 650 E. 25th Street
} Kansas City, MO 64108-2784
}
} Phone: (816) 235-2072
} Fax: (816) 235-5524
} Web: http://www.umkc.edu/dentistry/microscopy
}
}
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 16:33:19 2004



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Thu, 13 May 2004 16:39:38 -0500
Subject: [Microscopy] Sputter coater info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,
I would like to have an evaluation of the Denton Desk III TSC sputter
coater from those who are using or have used this instrument for preparation
of samples for FESEM imaging. Please direct your response to me rather than
to the list.

Many thanks,
Debby

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 18:42:36 2004



From: Gary Gaugler :      gary-at-gaugler.com
Date: Thu, 13 May 2004 16:48:04 -0700
Subject: [Microscopy] Re: graphite sputtering or coating

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I don't use C. I use Au/Pd and Pt. C is
too messy.

Depending on what you are looking at, either
image- or EDS/WDS-wise, your options may vary.
For EDS, I just ignore the small Au/Pd peaks.
The coating is only 40-60A and is typically
only seen as piled up Au with Al, Si and W.
With my specimens, Pd is by itself.

I suppose that if you were looking at P or Zr
specimens, Au/Pd or Pt would be a problem. Perhaps
the ultimate coating is Os. Someday, I would love
to try one of these coaters. It ought to be
great for EBSD.

But let us know what type of work you are trying
to do. That may make a big difference in coating
strategies.

gary g.


At 08:22 AM 5/13/2004, you wrote:


} hi all!
}
} a simple question :
}
} Does anyone have ever used graphite for sputtering or coating a SEM
} sample ?
}
} maybe a question from a dummy boy :-) , doesn't it?
}
} thanx for answers...
}
} Sylvain MAURY



From MicroscopyL-request-at-ns.microscopy.com Thu May 13 20:43:29 2004



From: Gary Gaugler :      gary-at-gaugler.com
Date: Thu, 13 May 2004 16:48:04 -0700
Subject: [Microscopy] Re: graphite sputtering or coating

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I don't use C. I use Au/Pd and Pt. C is
too messy.

Depending on what you are looking at, either
image- or EDS/WDS-wise, your options may vary.
For EDS, I just ignore the small Au/Pd peaks.
The coating is only 40-60A and is typically
only seen as piled up Au with Al, Si and W.
With my specimens, Pd is by itself.

I suppose that if you were looking at P or Zr
specimens, Au/Pd or Pt would be a problem. Perhaps
the ultimate coating is Os. Someday, I would love
to try one of these coaters. It ought to be
great for EBSD.

But let us know what type of work you are trying
to do. That may make a big difference in coating
strategies.

gary g.


At 08:22 AM 5/13/2004, you wrote:


} hi all!
}
} a simple question :
}
} Does anyone have ever used graphite for sputtering or coating a SEM
} sample ?
}
} maybe a question from a dummy boy :-) , doesn't it?
}
} thanx for answers...
}
} Sylvain MAURY




From MicroscopyL-request-at-ns.microscopy.com Fri May 14 08:16:15 2004



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Fri, 14 May 2004 10:15:09 -0400
Subject: [Microscopy] Re:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have an earlier model of the Epson (the 1600 Pro) and I"ve been
happy with it.
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Fri May 14 09:44:02 2004



From: Sherwood, Margaret :      MSHERWOOD-at-PARTNERS.ORG
Date: Fri, 14 May 2004 10:37:11 -0400
Subject: [Microscopy]

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Someone recently posed the same question on ListServer. We have an Epson
Perfection 3200 Photo and love it. There is no holder for the standard 3 1/4 X
4 negatives, but I just put them directly on the glass and they come out fine.

Peggy

Peggy Sherwood
Lab Associate, Photopathology
Wellman Laboratories of Photomedicine (W224)
Massachusetts General Hospital
55 Fruit Street
Boston, MA 02114
617-724-4839 (voice mail)
617-726-6983 (lab)
617-726-3192 (fax)
msherwood-at-partners.org


-----Original Message-----
} From: Dusevich, Vladimir [mailto:DusevichV-at-umkc.edu]
Sent: Thursday, May 13, 2004 2:26 PM
To: microscopy-at-msa.microscopy.com

I am going to buy scanner for TEM negatives and considering two models:
Microtek ArtixScan 1800f and Epson Expression 1680 Pro. Any advise and
comments are very welcome.

Thank you in advance,

Vladimir

Vladimir M. Dusevich, Ph.D.
Electron Microscope Lab Manager
3127 School of Dentistry
650 E. 25th Street
Kansas City, MO 64108-2784

Phone: (816) 235-2072
Fax: (816) 235-5524
Web: http://www.umkc.edu/dentistry/microscopy




From MicroscopyL-request-at-ns.microscopy.com Fri May 14 15:04:09 2004



From: AtcSEM :      atcsem-at-sbcglobal.net
Date: Fri, 14 May 2004 16:10:05 -0400
Subject: [Microscopy] IC chip cracking?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear microscopy experts I need your expertise.

We are evaluation transducer chips and suspecting that the chips have
pre-cracks forming between the side wall and the diaphragm. The problem: if
we do have a some sort of crack forming the gold coating, that I coat chips
for analysis, does not get inside the crack, so it's hard to prove if there
is any cracks or not. Do you have any suggestions how to show if there is
any cracks.

The arrows in the picture show the location of possible cracks.
http://www.atclabs.com/images/%231.jpg

Thanks,
Pavel Lozovyy
ATC SEM Lab
(216)692-6637
http://www.atclabs.com/SEM.htm




From MicroscopyL-request-at-ns.microscopy.com Fri May 14 16:30:39 2004



From: Gary Gaugler :      gary-at-gaugler.com
Date: Fri, 14 May 2004 14:36:05 -0700
Subject: [Microscopy] Re: IC chip cracking?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


What is the cross sectional construction like?
Is the die passivated? If it is, use MIL-STD-883 for
GLIT. If there is metal underneath the passivation,
and the passivation has voids, GLIT will show it
up big time. GLIT is glassivation layer integrity test.

gary g.


At 01:10 PM 5/14/2004, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Fri May 14 17:47:07 2004



From: Becky Holdford :      r-holdford-at-ti.com
Date: Fri, 14 May 2004 17:52:24 -0500
Subject: [Microscopy] Re: IC chip cracking?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Pavel: my first thought is, don't coat the device. Find an
accelerating voltage/tilt angle combination that will prevent of
minimize charging. Next thought, coat with something besides gold. But
I've done a lot of cracked devices and have never had any problems
telling the cracks from the original surface. Unless they are huge
cracks, the gold won't get in them to confuse the issue. Are these
surface cracks or cracks going down into the surface?

AtcSEM wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Becky Holdford (r-holdford-at-ti.com)
972-995-2360
972-648-8743 (pager)
SC Packaging FA Development
Texas Instruments, Inc.
Dallas, TX
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~




From MicroscopyL-request-at-ns.microscopy.com Sat May 15 00:47:25 2004



From: Alberto Diaspro :      diaspro-at-fisica.unige.it
Date: Sat, 15 May 2004 07:53:12 +0200
Subject: [Microscopy] Fluorescence School in Genoa at LAMBS on June 14-16 --- do not miss it!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
------------------------------------------------------------------------
------------------------
PRINCIPLES OF FLUORESCENCE TECHNIQUES COURSE 2004.

LAMBS, Department of Physics, University of Genoa, June 14-16, 2004.

The use and application of fluorescence techniques is increasing daily
both in the academia and in industry. The Principles of Fluorescence
Techniques course will outline the basic concepts of fluorescence
techniques and the successful utilization of the currently available
commercial instrumentation.  

The course is designed for students who utilize fluorescence
instrumentation and techniques, as well as for researchers
and industrial scientists who intend to deepen their knowledge of
fluorescence techniques. The theoretical lectures delivered by key
scientists in the field are complemented by the direct utilization of
steady state and lifetime fluorescence instrumentation provided by
leading companies.

It is recommended that participants have at least a bachelor's degree
in the life science, physical sciences or engineering before attending
this course. Topics addressed in this course include: 

Basic Definitions and Principles of Fluorescence
Fluorescence Polarization
Time- resolved Fluorescence
Instrumentation                                                         
                                                   
Data Manipulation and Data Analysis
Fiber Optic Sensors
Confocal Fluorescence Microscopy
Lifetime Imaging                                            
Fluorescence Correlation Spectroscopy
Multiphoton excitation microscopy and spectroscopy
Quantum dots 


 The Principles of Fluorescence Techniques course will be held at:

Laboratory for Advanced Microscopy, Bioimaging and Spectroscopy
Diaspro lab, Research Center MICROSCOBIO
Department of Physics, University of Genoa
via Dodecaneso, 33
16146 Genova, Italy

Details: www.lambs.it; diaspro-at-fisica.unige.it

The number of participants to the Principles of Fluorescence Techniques
Course

is limited to 40 people. PLEASE RESERVE YOUR PARTICIPATION AS SOON AS
POSSIBLE SENDING AN E-MAIL TO: coordinator-at-fluorescence-foundation.org.

Details : www.fluorescence-foundation.org
------------------------------------------------------------------------
------------------------------------------------------------------------
------------------
 





------------------------------------------------------------------------
---------------------------
Alberto Diaspro, Department of Physics, University of Genoa
Via Dodecaneso 33, 16146 Genova, Italy
facsimile +39-010314218 - voice +39-0103536426/480/309
URL: http://www.lambs.it
http://www.wiley.com/WileyCDA/WileyTitle/productCd-0471409200.html
------------------------------------------------------------------------
--------------------------




From MicroscopyL-request-at-ns.microscopy.com Sun May 16 13:24:22 2004



From: zr.zrzhang_xrli-at-ybb.ne.jp (by way of Ask-A-Microscopist)
Date: Sun, 16 May 2004 13:33:37 -0500
Subject: [Microscopy] AskAMicroscopist: TEM samples from Ag and Ag-Pd

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (zr zrzhang_xrli-at-ybb.ne.jp) from http://www.msa.microscopy.org/Ask-A-Microscopist.html on Saturday, May 15, 2004 at 21:16:24
---------------------------------------------------------------------------

Email: zr zrzhang_xrli-at-ybb.ne.jp
Name: Zheng-Rong Zhang

Organization: Yokohama National University

Education: Graduate College

Location: Yokohama, Japan

Question: Dear Microscopist,

I need to prepare TEM samples from Ag and Ag-Pd(10pct) alloy sheets. May you suggest me any electrolytes for twin-jet electropolishing and the corresponding conditions? No toxic chemicals would be used.

Thank you in advance!

With best regards

Zheng-Rong Zhang

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sun May 16 13:44:10 2004



From: jtrynak-at-SciGenium.com (by way of MicroscopyListserver)
Date: Sun, 16 May 2004 13:53:43 -0500
Subject: [Microscopy] viaWWW:Position Open for Sales Person

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (jtrynak-at-SciGenium.com) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, May 12, 2004 at 18:22:09
---------------------------------------------------------------------------

Email: jtrynak-at-SciGenium.com
Name: JT Rynak

Organization: SciGenium

Title-Subject: [Microscopy] [Filtered] Opening - New England Regional - Imaging Sales

Question: New England Regional - Imaging Sales

5+ years experience selling imaging / visualization capital equipment into Research Market space. Ideal person will have a rolodex of PI and Lead Vivo research scientists. BS/MS is Biology or Zoology

If you have any further questions feel free to contact me at 617-661-1067 or via email at jtrynak-at-scigenium.com


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sun May 16 18:53:06 2004



From: Tomic, Peter \(Peter\) :      ptomic-at-agere.com
Date: Sun, 16 May 2004 20:01:17 -0400
Subject: [Microscopy] IC chip cracking?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Pavel;

Having looked at more fractures in silicon and GaAs devices than I care to remember, it's not generally a problem finding them and as often with optical means as with EM.

If you would kindly provide the list with more specifics about your samples and the nature of these "cracks?" That is, are they along crystal planes or fractures that travel at random angles with respect to the surface? Fractures, or cleaves perpendicular to the die surface, are difficult to detect with EM but apparent with visible light microscopy. How separated are both sides of the crack and why do you need to see into them? Sounds like you only need to identify whether they exist or not. How long are the cracks?

I have plenty of EM's I'd be happy to send you images of samples that have fractured for various reasons, and there are multiple root causes of this failure mechanism. I'm sure if you provide a little more detail you'll get plenty of good suggestions on how to image your samples.

Regards,

Peter Tomic
Agere Systems

-----Original Message-----
} From: AtcSEM [mailto:atcsem-at-sbcglobal.net]
Sent: Friday, May 14, 2004 4:10 PM
To: MicroscopyListServer (MicroscopyListServer)

Dear microscopy experts I need your expertise.

We are evaluation transducer chips and suspecting that the chips have
pre-cracks forming between the side wall and the diaphragm. The problem: if
we do have a some sort of crack forming the gold coating, that I coat chips
for analysis, does not get inside the crack, so it's hard to prove if there
is any cracks or not. Do you have any suggestions how to show if there is
any cracks.

The arrows in the picture show the location of possible cracks.
http://www.atclabs.com/images/%231.jpg

Thanks,
Pavel Lozovyy
ATC SEM Lab
(216)692-6637
http://www.atclabs.com/SEM.htm






From MicroscopyL-request-at-ns.microscopy.com Sun May 16 19:26:15 2004



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Mon, 17 May 2004 12:37:11 +1200
Subject: [Microscopy] Re: Fluorescence School in Genoa at LAMBS on June 14-16 --- do not miss it!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Any course that will promote even the correct spelling of
fluorescence has to be a good thing.


Maybe someone will run one on how to spell 'JEOL' someday.

:)

cheers

rtch




From MicroscopyL-request-at-ns.microscopy.com Mon May 17 00:12:47 2004



From: Bryony James :      b.james-at-auckland.ac.nz
Date: Mon, 17 May 2004 17:21:33 +1200
Subject: [Microscopy] re: ESEM comparisons

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello listers,

We are in the process of justifying, to the University, a proposal to
purchase an ESEM. In prepapring the business case I am including the
ten year financials for both FEG and Tungtsen filament machines. Given
that we have to break even in that time, including covering our
depreciation, the Tungsten instrument is easier to justify in a purely
financial sense.

From a scientific and operational viewpoint, though, I was hoping some
of you could provide opinions on the relative merits of Tungsten or FEG.
Our current conventional SEM is a Philips XL-30 S-FEG so we already know
of the costs associated with the emitter.

The range of samples we will be required to put in our ESEM will be
anything you oculd possibly imagine from the Faculties of Engineering,
Science and Medicine. Samples will inlcude food materials (big dairy
industry in New Zealand), polymers, conducting polymers, mammalian
tissue, bacteria, fruit and veg, fish and chips, superconductors,
ceramic coatings and bees knees. We will definitely need a hot-stage.

If you would like to contact me off list with your opinions please feel
free.

Kind regards
Bryony James
Research Centre for Surface and Materials Science
Univetrsity of Auckland
New Zealand


From MicroscopyL-request-at-ns.microscopy.com Mon May 17 07:57:52 2004



From: AtcSEM :      atcsem-at-sbcglobal.net
Date: Mon, 17 May 2004 09:06:52 -0400
Subject: [Microscopy] IC chip cracking?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thank you all for the responses that I have received.

Just to clarify:
We had multiple field failures of the chips. The SEM examination
revealed some sort of coating on the fracture surfaces, which I believe is
the urethane coating. The transition from the side wall to the fracture
surface, on some of the chips, suggests that the chips could have had a
small cracks prior urethane coating.

Now, I am looking at the "Virgin" chips for any evidence of cracking
at the side wall and diaphragm transition, but my problem is that the Gold
coating does not coat the area well enough for me to resolve the crack. Due
to lack of coating that area is charging and I see only a black line.
I have coated it at least 3 times and rotate the chips to insure good
coatings in those areas, but it was not enough.

After SEM examination I would also like to cross section the chip to check
for the evidence of pre-cracks, but I am afraid to introduce the cracks
myself. Could you help me with a proper preparation procedure of the cross
sections? We have metallography lab. equipment and nothing special to
prepare the chip cross sections.

I have uploaded some more photos of the fracture surfaces of the failed
chips and Virgin chip that I am trying to image now.
http://www.atclabs.com/images/temp/MicrList_files/frame.htm

Sorry, it might be a little slow at showing the images due to their size.

Thanks for your help again,

Pavel


-----Original Message-----
} From: Gary Gaugler [mailto:gary-at-gaugler.com]

What is the cross sectional construction like?
Is the die passivated? If it is, use MIL-STD-883 for
GLIT. If there is metal underneath the passivation,
and the passivation has voids, GLIT will show it
up big time. GLIT is glassivation layer integrity test.

gary g.

-----Original Message-----
} From: Warren E Straszheim [mailto:wesaia-at-iastate.edu]

Do you get charging at the crack due to a gap in the coating?
Otherwise,you should be able to see the break.

-----Original Message-----
} From: Jim Quinn [mailto:jquinn-at-www.matscieng.sunysb.edu]
Pavel

Your image is very low mag, so it is difficult to judge anything.

I suggest that you use a sacrificial chip and coat
it with about 10nm of gold. If your coater does not
rotate the sample, then you should manually rotate
the sample at 45 degrees atleast twice. This is
will help the gold to get into the perpendicular
surfaces.

Next use a Robinson BS detector. 10 or 15kV is sufficient.

Once you image the cracks, then you can go about
trying to image without the gold. Variable pressure
and voltage will reduce charging in uncoated samples.

JQ


-----Original Message-----
} From: Tomic, Peter (Peter) [mailto:ptomic-at-agere.com]
Pavel;

Having looked at more fractures in silicon and GaAs devices than I
care to remember, it's not generally a problem finding them and as
often with optical means as with EM.

If you would kindly provide the list with more specifics about your
samples and the nature of these "cracks?" That is, are they along
crystal planes or fractures that travel at random angles with respect to
the surface? Fractures, or cleaves perpendicular to the die surface, are
difficult to detect with EM but apparent with visible light microscopy.
How separated are both sides of the crack and why do you need to see into
them? Sounds like you only need to identify whether they exist or not.
How long are the cracks?

I have plenty of EM's I'd be happy to send you images of samples
that have fractured for various reasons, and there are multiple root
causes of this failure mechanism. I'm sure if you provide a little more
detail you'll get plenty of good suggestions on how to image your
samples.

Regards,

Peter Tomic
Agere Systems

-----Original Message-----
} From: AtcSEM [mailto:atcsem-at-sbcglobal.net]
Sent: Friday, May 14, 2004 4:10 PM
To: MicroscopyListServer (MicroscopyListServer)

Dear microscopy experts I need your expertise.

We are evaluation transducer chips and suspecting that the chips have
pre-cracks forming between the side wall and the diaphragm. The problem: if
we do have a some sort of crack forming the gold coating, that I coat chips
for analysis, does not get inside the crack, so it's hard to prove if there
is any cracks or not. Do you have any suggestions how to show if there is
any cracks.

The arrows in the picture show the location of possible cracks.
http://www.atclabs.com/images/%231.jpg

Thanks,
Pavel Lozovyy
ATC SEM Lab
(216)692-6637
http://www.atclabs.com/SEM.htm





From MicroscopyL-request-at-ns.microscopy.com Mon May 17 09:11:57 2004



From: sghoshro-at-NMSU.Edu
Date: Mon, 17 May 2004 08:21:10 -0600 (MDT)
Subject: [Microscopy] GFP survival, thank you

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thank you so much for all of your advice. I received a large number of
great suggestions and almost everyone suggested using anti GFP antibody
and we are certainly going to go that route.

Thanks again and this is truly a great resource.

Best wishes,

Soumitra

*************************************************************
Soumitra Ghoshroy
College Associate Professor, Biology
Director, Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
http://www.cs.nmsu.edu/~biology/Faculty%20&%20Staff/Ghoshroy/Ghoshroy.htm
http://emldata.nmsu.edu/eml/


From MicroscopyL-request-at-ns.microscopy.com Mon May 17 11:36:59 2004



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Mon, 17 May 2004 11:43:51 -0500
Subject: [Microscopy] Re: RE: IC chip cracking?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I think I can offer a few ideas and questions.

1) Can you say "stress concentration"? My academic background is in
Engineering Science and Mechanics. It looks like you have some very sharp
corners where the diaphragm meets the substrate. That is trouble waiting to
happen. I cannot tell if there is any pre-crack there. There might be hints
of one in slide 43. Either way, that geometry is going to lead to problems.
The stresses will be highest at the edges of the diaphragm. I think you
would prefer a gradual change in geometry there to avoid any concentration
and to help reinforce the edges. I don't know the fabrication process, and
it might be hard to implement, but it might be worth investigating.

Slides 38 and 42 show some jagged surfaces. I guess those are the edges of
the substrate and the diaphragm is on the bottom. If so, the nose of the
jaggie jutting out further into the opening will be a point of stress
concentration.

You may also want to consider round openings in the substrate. The corners
of the square holes don't help and might hurt. The highest stresses should
be in the middle of the edges.

2) You haven't said how prematurely these chips failed. Are they subjected
to a cyclic load? What fraction of their design load were they subjected
to? There may be hints of fatigue failure (slide 34?), but it seems to be
rather short cycle fatigue if it is fatigue at all. I have seen glass break
with such fractures in overload.

3) Slides 37, 40 and 41 show dark corners. I think that is the normal
appearance for an interior corner in secondary mode. An exterior corner
tends to show up bright in SE mode.

4) I can't quite tell where many of the photos were taken (e.g., slides 13
and 39). I usually use a three-fold step in magnification to easily follow
from one view to another. Also, the progressions 10, 30, 100, 300 or 15,
50, 150, 500 are easy to remember and to standardize on.

5) Did you copy these images into Powerpoint to make this presentation? My
computer took quite a while to load each slide, and I have a fast
connection. We routinely record our SEM images at 1024-pixels across and
store the files as JPGs. (Keep reading for the rationale.) Those files are
maybe 300 KB in size. If I use the Insert, Image, From_file function to add
the image to my document, it retains the JPG compression and gray nature
and requires only 300 KB. But, if I open those files in an imaging program
and copy them over they expand. The computer has the image rendered as a
bitmap in full color, the image gets copied with no compression and it now
consumes 2400 kB (1024x768 pixels times 3-bytes-per-pixel). That 8-fold
difference in size is noticeable. I know more and more users have broadband
connections; however these order-of-magnitude expansions start to add up. {g}

Hope this helps.

Warren

At 08:06 AM 5/17/2004, you wrote:

} Thank you all for the responses that I have received.
}
} Just to clarify:
} We had multiple field failures of the chips. The SEM
} examination
} revealed some sort of coating on the fracture surfaces, which I believe is
} the urethane coating. The transition from the side wall to the fracture
} surface, on some of the chips, suggests that the chips could have had a
} small cracks prior urethane coating.
}
} Now, I am looking at the "Virgin" chips for any evidence of cracking
} at the side wall and diaphragm transition, but my problem is that the Gold
} coating does not coat the area well enough for me to resolve the crack. Due
} to lack of coating that area is charging and I see only a black line.
} I have coated it at least 3 times and rotate the chips to insure good
} coatings in those areas, but it was not enough.
}
} After SEM examination I would also like to cross section the chip to check
} for the evidence of pre-cracks, but I am afraid to introduce the cracks
} myself. Could you help me with a proper preparation procedure of the cross
} sections? We have metallography lab. equipment and nothing special to
} prepare the chip cross sections.
}
} I have uploaded some more photos of the fracture surfaces of the failed
} chips and Virgin chip that I am trying to image now.
} http://www.atclabs.com/images/temp/MicrList_files/frame.htm
}
} Sorry, it might be a little slow at showing the images due to their size.
}
} Thanks for your help again,
}
} Pavel

-------------------------------------------
No files should be attached to this message
-------------------------------------------
Warren E. Straszheim, Ph.D.
Materials Analysis and Research Lab
Iowa State University
46 Town Engineering
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
Web: www.marl.iastate.edu

Scanning electron microscopy, x-ray analysis, and image analysis of materials
Computer applications and networking



From MicroscopyL-request-at-ns.microscopy.com Mon May 17 11:44:19 2004



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Mon, 17 May 2004 17:54:03 +0100 (GMT Daylight Time)
Subject: [Microscopy] Re: re: ESEM comparisons

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We look mostly at hydrated biological samples. We have a
tungsten gun XL30 ESEM TMP. For looking at bacteria and
cells a FEGESEM would be superior in terms of reduced
damage and improved resolution. I find it hard to get
medium/high resolution with wet biological samples. I feel
FEGESEM would make this a lot easier.

Dave

On Mon, 17 May 2004 17:21:33 +1200 Bryony James
{b.james-at-auckland.ac.nz} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Hello listers,
}
} We are in the process of justifying, to the University, a proposal to
} purchase an ESEM. In prepapring the business case I am including the
} ten year financials for both FEG and Tungtsen filament machines. Given
} that we have to break even in that time, including covering our
} depreciation, the Tungsten instrument is easier to justify in a purely
} financial sense.
}
} From a scientific and operational viewpoint, though, I was hoping some
} of you could provide opinions on the relative merits of Tungsten or FEG.
} Our current conventional SEM is a Philips XL-30 S-FEG so we already know
} of the costs associated with the emitter.
}
} The range of samples we will be required to put in our ESEM will be
} anything you oculd possibly imagine from the Faculties of Engineering,
} Science and Medicine. Samples will inlcude food materials (big dairy
} industry in New Zealand), polymers, conducting polymers, mammalian
} tissue, bacteria, fruit and veg, fish and chips, superconductors,
} ceramic coatings and bees knees. We will definitely need a hot-stage.
}
} If you would like to contact me off list with your opinions please feel
} free.
}
} Kind regards
} Bryony James
} Research Centre for Surface and Materials Science
} Univetrsity of Auckland
} New Zealand
}
}
}
} This incoming email to UWE has been independently scanned for viruses and any virus detected has been removed using McAfee anti-virus software
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



This email has been independently scanned for viruses and any virus detected has been removed using McAfee anti-virus software



From MicroscopyL-request-at-ns.microscopy.com Mon May 17 13:37:31 2004



From: Alan Stone :      as-at-astonmet.com
Date: Mon, 17 May 2004 13:46:45 -0500
Subject: [Microscopy] Legal/Ethics Safety Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


We are an independent commercial lab serving general industry. Unless
directed by a court order or we suspect misuse of our findings, all
information is strictly confidential to our client.

We have a situation with a small project we just completed. We detected the
presence of a toxic compound in product. While not the intended use, it is
conceivable that this product could be put into one's mouth which would
present a health concern.

Upon discovery of this compound, I called our client immediately and found
out that they submitted competitors' products. Our client is certainly
aware of this now, but they are not in a position to disclose to their
competitors that they were analyzing their products. Given that these
products are not intended to have oral contact, I doubt that the FDA has
any applicable regulations regarding the composition.

Any thoughts?

Alan Stone
ASTON Metallurgical Services






Alan Stone
ASTON Metallurgical Services Co., Inc.
200 Larkin Drive Ste A
Wheeling, IL 60090
847/353-8100
www.astonmet.com



From MicroscopyL-request-at-ns.microscopy.com Mon May 17 16:10:25 2004



From: Tomic, Peter \(Peter\) :      ptomic-at-agere.com
Date: Tue, 18 May 2004 08:05:28 -0400
Subject: [Microscopy] Legal/Ethics Safety Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Bryony,
The combined FE and ESED instruments are very pricey, expensive to maintain and,
since you already have a FE, perhaps a bit redundant. We have a two-year-old
tungsten variable-pressure and it is a popular workhorse that is in constant
use. It can do high resolution or ESED but not both at the same time. If you
need both at the same time, then you need FE-ESED. BTW, the thing we would most
like to have is a cold stage, to look at hydrated samples without having to go
to higher pressures.
Good luck.
Regards,
Mary Mager
Electron Microscopist
Department of Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchange.ubc.ca
----- Original Message -----
} From: "Bryony James" {b.james-at-auckland.ac.nz}
To: {Microscopy-at-msa.microscopy.com}
Sent: Sunday, May 16, 2004 10:21 PM

Alan;

It sounds like you need legal advice and not scientific advice. You should consult with your corporate counsel on this issue.

Competitor analysis is done all the time and I don't see any issues with that. What you do with the data afterwards may be another matter and open to litigation so see the attorney.

Regards,

Peter Tomic
Agere Systems

-----Original Message-----
} From: Alan Stone [mailto:as-at-astonmet.com]
Sent: Monday, May 17, 2004 2:47 PM
To: Microscopy-at-MSA.Microscopy.Com


We are an independent commercial lab serving general industry. Unless
directed by a court order or we suspect misuse of our findings, all
information is strictly confidential to our client.

We have a situation with a small project we just completed. We detected the
presence of a toxic compound in product. While not the intended use, it is
conceivable that this product could be put into one's mouth which would
present a health concern.

Upon discovery of this compound, I called our client immediately and found
out that they submitted competitors' products. Our client is certainly
aware of this now, but they are not in a position to disclose to their
competitors that they were analyzing their products. Given that these
products are not intended to have oral contact, I doubt that the FDA has
any applicable regulations regarding the composition.

Any thoughts?

Alan Stone
ASTON Metallurgical Services






Alan Stone
ASTON Metallurgical Services Co., Inc.
200 Larkin Drive Ste A
Wheeling, IL 60090
847/353-8100
www.astonmet.com





From MicroscopyL-request-at-ns.microscopy.com Tue May 18 07:02:20 2004



From: Tomic, Peter \(Peter\) :      ptomic-at-agere.com
Date: Tue, 18 May 2004 08:10:31 -0400
Subject: [Microscopy] IC chip cracking?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Pavel;

If you really need a x-section without the issues with mechanical polishing, try using an FIB [Focused Ion Beam].

Peter

-----Original Message-----
} From: AtcSEM [mailto:atcsem-at-sbcglobal.net]
Sent: Monday, May 17, 2004 9:07 AM
To: MicroscopyListServer (MicroscopyListServer)

Thank you all for the responses that I have received.

Just to clarify:
We had multiple field failures of the chips. The SEM examination
revealed some sort of coating on the fracture surfaces, which I believe is
the urethane coating. The transition from the side wall to the fracture
surface, on some of the chips, suggests that the chips could have had a
small cracks prior urethane coating.

Now, I am looking at the "Virgin" chips for any evidence of cracking
at the side wall and diaphragm transition, but my problem is that the Gold
coating does not coat the area well enough for me to resolve the crack. Due
to lack of coating that area is charging and I see only a black line.
I have coated it at least 3 times and rotate the chips to insure good
coatings in those areas, but it was not enough.

After SEM examination I would also like to cross section the chip to check
for the evidence of pre-cracks, but I am afraid to introduce the cracks
myself. Could you help me with a proper preparation procedure of the cross
sections? We have metallography lab. equipment and nothing special to
prepare the chip cross sections.

I have uploaded some more photos of the fracture surfaces of the failed
chips and Virgin chip that I am trying to image now.
http://www.atclabs.com/images/temp/MicrList_files/frame.htm

Sorry, it might be a little slow at showing the images due to their size.

Thanks for your help again,

Pavel


-----Original Message-----
} From: Gary Gaugler [mailto:gary-at-gaugler.com]

What is the cross sectional construction like?
Is the die passivated? If it is, use MIL-STD-883 for
GLIT. If there is metal underneath the passivation,
and the passivation has voids, GLIT will show it
up big time. GLIT is glassivation layer integrity test.

gary g.

-----Original Message-----
} From: Warren E Straszheim [mailto:wesaia-at-iastate.edu]

Do you get charging at the crack due to a gap in the coating?
Otherwise,you should be able to see the break.

-----Original Message-----
} From: Jim Quinn [mailto:jquinn-at-www.matscieng.sunysb.edu]
Pavel

Your image is very low mag, so it is difficult to judge anything.

I suggest that you use a sacrificial chip and coat
it with about 10nm of gold. If your coater does not
rotate the sample, then you should manually rotate
the sample at 45 degrees atleast twice. This is
will help the gold to get into the perpendicular
surfaces.

Next use a Robinson BS detector. 10 or 15kV is sufficient.

Once you image the cracks, then you can go about
trying to image without the gold. Variable pressure
and voltage will reduce charging in uncoated samples.

JQ


-----Original Message-----
} From: Tomic, Peter (Peter) [mailto:ptomic-at-agere.com]
Pavel;

Having looked at more fractures in silicon and GaAs devices than I
care to remember, it's not generally a problem finding them and as
often with optical means as with EM.

If you would kindly provide the list with more specifics about your
samples and the nature of these "cracks?" That is, are they along
crystal planes or fractures that travel at random angles with respect to
the surface? Fractures, or cleaves perpendicular to the die surface, are
difficult to detect with EM but apparent with visible light microscopy.
How separated are both sides of the crack and why do you need to see into
them? Sounds like you only need to identify whether they exist or not.
How long are the cracks?

I have plenty of EM's I'd be happy to send you images of samples
that have fractured for various reasons, and there are multiple root
causes of this failure mechanism. I'm sure if you provide a little more
detail you'll get plenty of good suggestions on how to image your
samples.

Regards,

Peter Tomic
Agere Systems

-----Original Message-----
} From: AtcSEM [mailto:atcsem-at-sbcglobal.net]
Sent: Friday, May 14, 2004 4:10 PM
To: MicroscopyListServer (MicroscopyListServer)

Dear microscopy experts I need your expertise.

We are evaluation transducer chips and suspecting that the chips have
pre-cracks forming between the side wall and the diaphragm. The problem: if
we do have a some sort of crack forming the gold coating, that I coat chips
for analysis, does not get inside the crack, so it's hard to prove if there
is any cracks or not. Do you have any suggestions how to show if there is
any cracks.

The arrows in the picture show the location of possible cracks.
http://www.atclabs.com/images/%231.jpg

Thanks,
Pavel Lozovyy
ATC SEM Lab
(216)692-6637
http://www.atclabs.com/SEM.htm







From MicroscopyL-request-at-ns.microscopy.com Tue May 18 07:15:34 2004



From: Claude Grattepain :      claude.grattepain-at-thalesgroup.com
Date: Tue, 18 May 2004 14:24:29 +0200
Subject: [Microscopy] SEM_need help on HfO2/Si imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello,
We want to image, by SEM, the surface of a HfO2 layer deposited on Si.
Is there a method to reveal the structure of the layer ?
for instance chemical etching or others....
We have tried on the as grown layer but we can see nothing, the image is
very smooth. By means of X-ray diffraction we know that
we have very small crystals, just a few nanometers. Now we wonder wether
there are larger structures than the crystals. We expect
something like polycrystalline grains...
Thank you for your help
Regards

--
Claude GRATTEPAIN / Bernard SERVET

THALES R & T France Tel : (+33 1) 69.33.91.13
Domaine de Corbeville Fax : (+33 1) 69.33.07.40

91404 Orsay Cedex, France mailto:claude.grattepain-at-thalesgroup.com
_____________________________________________________________________

"The information contained in this e-mail and any attachments are the
property of THALES and may be confidential. If you are not the
intended
recipient, please notify us immediately, send this message back to us
and
destroy it. You are hereby notified that any review, dissemination,
distribution, copying or otherwise use of this e-mail/fax is strictly
prohibited.




From MicroscopyL-request-at-ns.microscopy.com Tue May 18 08:17:11 2004



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Tue, 18 May 2004 15:30:40 +0200
Subject: [Microscopy] [microscopy] Centrifuge for Virus separation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear All

We would like to get a centrifuge to separate viruses. What are you using in your TEM laboratory?
Suggestions and hints will be helpful.
Thanks

Since my UB mail address is not reliable please send
a copy of all urgent mail to

coetzeesh-at-yahoo.co.uk

Mr S. H. Coetzee
Electron Microscope Unit
University of Botswana
Private Bag 0022
Gaborone
Botswana
Phone : +267 355 2462/5222
Mobile : +267 714 55701 (Now active)
Fax : +267 318 5097
e-mail : coetzees-at-mopipi.ub.bw



From MicroscopyL-request-at-ns.microscopy.com Tue May 18 08:46:08 2004



From: Lee, Steven :      StevenLe-at-BaylorHealth.edu
Date: Tue, 18 May 2004 08:55:24 -0500
Subject: [Microscopy] TEM Film scanner Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html




Hello Listers,

I wanted to see if anyone has had any experience scanning negatives with the Canon Canoscan 9900F. 
I have a low end Canoscan that I use for documents and the occasional photograph, which has given me
very good results, but is not set up for negatives.  I have seen the recent posts about the Epson scanner,
but I am trying to keep it under a $1000 so that I do not have to treat it as a capital purchase.  I have tried
for the last two years to get a Coolscan 8000 on the capital budget with no luck, and do not see any scanner
above $1000 as an option right now.

Thanks,
Steve


Steven Lee
Chief Technologists
Electron Microscopy Laboratory
Baylor University Medical Center
3500 Gaston Ave
Dallas, TX 750246
Ph: 214.820.3302
Fx: 214.820.4110
Em: stevenle-at-baylorhealth.edu



This e-mail, facsimile, or letter and any files or attachments transmitted with it contains information that is confidential and privileged. This information is intended only for the use of the individual(s) and entity(ies) to whom it is addressed. If you are the intended recipient, further disclosures are prohibited without proper authorization. If you are not the intended recipient, any disclosure, copying, printing, or use of this information is strictly prohibited and possibly a violation of federal or state law and regulations. If you have received this information in error, please notify Baylor Health Care System immediately at 1-866-402-1661 or via e-mail at privacy-at-baylorhealth.edu. Baylor Health Care System, its subsidiaries, and affiliates hereby claim all applicable privileges related to this information.



From MicroscopyL-request-at-ns.microscopy.com Tue May 18 08:47:42 2004



From: Dusevich, Vladimir :      DusevichV-at-umkc.edu
Date: Tue, 18 May 2004 08:55:19 -0500
Subject: [Microscopy] RE: Re: re: ESEM comparisons

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

In my experience medium/high magnifications for wet
specimens are limited mostly by specimens themselves and not by a
type of emitter. While I can get a nice picture of gold on carbon
at x100k, for most of my "real life" specimens magnification is
limited by x20k (or less).

Vladimir

} We look mostly at hydrated biological samples. We have a
} tungsten gun XL30 ESEM TMP. For looking at bacteria and
} cells a FEGESEM would be superior in terms of reduced
} damage and improved resolution. I find it hard to get
} medium/high resolution with wet biological samples. I feel
} FEGESEM would make this a lot easier.
}
} Dave
}
} On Mon, 17 May 2004 17:21:33 +1200 Bryony James
} {b.james-at-auckland.ac.nz} wrote:
}
} }
} }
} }
} ----------------------------------------------------------------------
} } --------
} } The Microscopy ListServer -- Sponsor: The Microscopy
} Society of America
} } To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} --------------------------------------------------------------
} -----------------
} }
} } Hello listers,
} }
} } We are in the process of justifying, to the University, a
} proposal to
} } purchase an ESEM. In prepapring the business case I am
} including the
} } ten year financials for both FEG and Tungtsen filament
} machines. Given
} } that we have to break even in that time, including covering our
} } depreciation, the Tungsten instrument is easier to justify
} in a purely
} } financial sense.
} }
} } From a scientific and operational viewpoint, though, I was hoping
} } some
} } of you could provide opinions on the relative merits of
} Tungsten or FEG.
} } Our current conventional SEM is a Philips XL-30 S-FEG so we
} already know
} } of the costs associated with the emitter.
} }
} } The range of samples we will be required to put in our ESEM will be
} } anything you oculd possibly imagine from the Faculties of
} Engineering,
} } Science and Medicine. Samples will inlcude food materials
} (big dairy
} } industry in New Zealand), polymers, conducting polymers, mammalian
} } tissue, bacteria, fruit and veg, fish and chips, superconductors,
} } ceramic coatings and bees knees. We will definitely need a
} hot-stage.
} }
} } If you would like to contact me off list with your opinions please
} } feel
} } free.
} }
} } Kind regards
} } Bryony James
} } Research Centre for Surface and Materials Science
} } Univetrsity of Auckland
} } New Zealand
} }
} }
} }
} } This incoming email to UWE has been independently scanned
} for viruses
} } and any virus detected has been removed using McAfee anti-virus
} } software
} }
}
} ----------------------------------------
} Patton, David
} Email: David.Patton-at-uwe.ac.uk
} "University of the West of England"
}
}
}
} This email has been independently scanned for viruses and any
} virus detected has been removed using McAfee anti-virus software
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Tue May 18 08:51:20 2004



From: Gary Gaugler :      gary-at-gaugler.com
Date: Tue, 18 May 2004 06:59:52 -0700
Subject: [Microscopy] Re: SEM_need help on HfO2/Si imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Try EBSD. It will tell if you have grains
or if the layer is amorphous.

gary g.


At 05:24 AM 5/18/2004, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Tue May 18 09:16:24 2004



From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Tue, 18 May 2004 15:25:31 +0100 (GMT Daylight Time)
Subject: [Microscopy] RE: Re: re: ESEM comparisons

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Depressing isn't it? I usually give up with cultured
cells and look at them after drying with HMDS and gold
coating. Have you (or anyone out there) ever taken/seen a
good picture of a hydrated cultured cell at x20,000 or
higher?

Dave

On Tue, 18 May 2004 08:55:19 -0500 "Dusevich, Vladimir"
{DusevichV-at-umkc.edu} wrote:

} In my experience medium/high magnifications for wet
} specimens are limited mostly by specimens themselves and not by a
} type of emitter. While I can get a nice picture of gold on carbon
} at x100k, for most of my "real life" specimens magnification is
} limited by x20k (or less).
}
} Vladimir
}
} } We look mostly at hydrated biological samples. We have a
} } tungsten gun XL30 ESEM TMP. For looking at bacteria and
} } cells a FEGESEM would be superior in terms of reduced
} } damage and improved resolution. I find it hard to get
} } medium/high resolution with wet biological samples. I feel
} } FEGESEM would make this a lot easier.
} }
} } Dave
} }
} } On Mon, 17 May 2004 17:21:33 +1200 Bryony James
} } {b.james-at-auckland.ac.nz} wrote:
} }
} } }
} } }
} } }
} } ----------------------------------------------------------------------
} } } --------
} } } The Microscopy ListServer -- Sponsor: The Microscopy
} } Society of America
} } } To Subscribe/Unsubscribe --
} } http://www.msa.microscopy.com/MicroscopyListserver
} } } On-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} } --------------------------------------------------------------
} } -----------------
} } }
} } } Hello listers,
} } }
} } } We are in the process of justifying, to the University, a
} } proposal to
} } } purchase an ESEM. In prepapring the business case I am
} } including the
} } } ten year financials for both FEG and Tungtsen filament
} } machines. Given
} } } that we have to break even in that time, including covering our
} } } depreciation, the Tungsten instrument is easier to justify
} } in a purely
} } } financial sense.
} } }
} } } From a scientific and operational viewpoint, though, I was hoping
} } } some
} } } of you could provide opinions on the relative merits of
} } Tungsten or FEG.
} } } Our current conventional SEM is a Philips XL-30 S-FEG so we
} } already know
} } } of the costs associated with the emitter.
} } }
} } } The range of samples we will be required to put in our ESEM will be
} } } anything you oculd possibly imagine from the Faculties of
} } Engineering,
} } } Science and Medicine. Samples will inlcude food materials
} } (big dairy
} } } industry in New Zealand), polymers, conducting polymers, mammalian
} } } tissue, bacteria, fruit and veg, fish and chips, superconductors,
} } } ceramic coatings and bees knees. We will definitely need a
} } hot-stage.
} } }
} } } If you would like to contact me off list with your opinions please
} } } feel
} } } free.
} } }
} } } Kind regards
} } } Bryony James
} } } Research Centre for Surface and Materials Science
} } } Univetrsity of Auckland
} } } New Zealand
} } }
} } }
} } }
} } } This incoming email to UWE has been independently scanned
} } for viruses
} } } and any virus detected has been removed using McAfee anti-virus
} } } software
} } }
} }
} } ----------------------------------------
} } Patton, David
} } Email: David.Patton-at-uwe.ac.uk
} } "University of the West of England"
} }
} }
} }
} } This email has been independently scanned for viruses and any
} } virus detected has been removed using McAfee anti-virus software
} }
} }
} }
}
}
} This incoming email to UWE has been independently scanned for viruses and any virus detected has been removed using McAfee anti-virus software
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"



This email has been independently scanned for viruses and any virus detected has been removed using McAfee anti-virus software



From MicroscopyL-request-at-ns.microscopy.com Tue May 18 10:42:01 2004



From: Tom Phillips :      phillipst-at-missouri.edu
Date: Tue, 18 May 2004 10:50:33 -0500
Subject: [Microscopy] Re: TEM Film scanner Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

When choosing a scanner for negatives, an important parameter is the Dmax
(which is a log value so relatively small differences can be
significant). The Dmax needed for negs is a lot more than for most printed
images or documents.

At 08:55 AM 5/18/2004 -0500, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thomas E. Phillips, PhD
Professor of Biological Sciences
Director, Molecular Cytology Core
3 Tucker Hall
University of Missouri
Columbia, MO 65211-7400

573-882-4712 (office)
573-882-0123 (fax)
PhillipsT-at-missouri.edu




From MicroscopyL-request-at-ns.microscopy.com Tue May 18 11:16:44 2004



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Tue, 18 May 2004 11:25:59 -0500
Subject: [Microscopy] Re: RE: Re: re: ESEM comparisons

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I agree that resolution is more dependent on sample than microscope. What a
microscope is capable of doing with an "ideal" sample often does not relate
to the real world of biological samples which are anything but ideal!

To get really good resolution you need to use higher kV values. These same
values will result in the primary electrons penetrating too far into the low
density biological sample resulting in substantial loss of signal in general
and se1 & se2's in particular. Resulting image will be low resolution and
often of much lesser quality than the same sample taken at a lower
magnification.

The problem of kV can be partially overcome by using a more coherent
beam with greater beam density, i.e. Field Emission. FE allows one to work
at lower kV values resulting in a se yield that is still adequate while
minimizing sample damage. Resulting images are likely to be higher
resolution (sharper with more detail) than comparable images at similar kVs
from a tungsten instrument.

However, there is a point where you go from useful magnification to
"empty " magnification regardless of the gun type. This is sample dependent
and must be determined accordingly. So it is unlikely that you will get any
additional information from a bacteria sample at 100k than at say 50K.
However you will most likely get a much sharper image at 20k with FE than
you could with a W filament.

Debby



Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy


On 5/18/04 8:55 AM, "Dusevich, Vladimir" {DusevichV-at-umkc.edu} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
----------------------------------------------------------------------------
--}
-
}
} In my experience medium/high magnifications for wet
} specimens are limited mostly by specimens themselves and not by a
} type of emitter. While I can get a nice picture of gold on carbon
} at x100k, for most of my "real life" specimens magnification is
} limited by x20k (or less).
}
} Vladimir
}
} } We look mostly at hydrated biological samples. We have a
} } tungsten gun XL30 ESEM TMP. For looking at bacteria and
} } cells a FEGESEM would be superior in terms of reduced
} } damage and improved resolution. I find it hard to get
} } medium/high resolution with wet biological samples. I feel
} } FEGESEM would make this a lot easier.
} }
} } Dave
} }
} } On Mon, 17 May 2004 17:21:33 +1200 Bryony James
} } {b.james-at-auckland.ac.nz} wrote:
} }
} } }
} } }
} } }
} } ----------------------------------------------------------------------
} } } --------
} } } The Microscopy ListServer -- Sponsor: The Microscopy
} } Society of America
} } } To Subscribe/Unsubscribe --
} } http://www.msa.microscopy.com/MicroscopyListserver
} } } On-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} } --------------------------------------------------------------
} } -----------------
} } }
} } } Hello listers,
} } }
} } } We are in the process of justifying, to the University, a
} } proposal to
} } } purchase an ESEM. In prepapring the business case I am
} } including the
} } } ten year financials for both FEG and Tungtsen filament
} } machines. Given
} } } that we have to break even in that time, including covering our
} } } depreciation, the Tungsten instrument is easier to justify
} } in a purely
} } } financial sense.
} } }
} } } From a scientific and operational viewpoint, though, I was hoping
} } } some
} } } of you could provide opinions on the relative merits of
} } Tungsten or FEG.
} } } Our current conventional SEM is a Philips XL-30 S-FEG so we
} } already know
} } } of the costs associated with the emitter.
} } }
} } } The range of samples we will be required to put in our ESEM will be
} } } anything you oculd possibly imagine from the Faculties of
} } Engineering,
} } } Science and Medicine. Samples will inlcude food materials
} } (big dairy
} } } industry in New Zealand), polymers, conducting polymers, mammalian
} } } tissue, bacteria, fruit and veg, fish and chips, superconductors,
} } } ceramic coatings and bees knees. We will definitely need a
} } hot-stage.
} } }
} } } If you would like to contact me off list with your opinions please
} } } feel
} } } free.
} } }
} } } Kind regards
} } } Bryony James
} } } Research Centre for Surface and Materials Science
} } } Univetrsity of Auckland
} } } New Zealand
} } }
} } }
} } }
} } } This incoming email to UWE has been independently scanned
} } for viruses
} } } and any virus detected has been removed using McAfee anti-virus
} } } software
} } }
} }
} } ----------------------------------------
} } Patton, David
} } Email: David.Patton-at-uwe.ac.uk
} } "University of the West of England"
} }
} }
} }
} } This email has been independently scanned for viruses and any
} } virus detected has been removed using McAfee anti-virus software
} }
} }
} }
}
}
}





From MicroscopyL-request-at-ns.microscopy.com Tue May 18 11:19:50 2004



From: Young, Gene (GP) :      GPYoung-at-dow.com
Date: Tue, 18 May 2004 11:42:45 -0400
Subject: [Microscopy] RE: TEM Film scanner Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I haven't tried the scanners you mentioned but you may want to check out this April 2004 review article in PCWorld online:

http://www.pcworld.com/reviews/article/0,aid,115075,00.asp

Gene P. Young
Sr. Analytical Technologist
Analytical Sciences, Polymer Characterization

The Dow Chemical Company
2301 N. Brazosport Blvd., B-1470
Freeport, Texas 77541-3257
Fax: (979) 238-0095

*(979) 238-1579 * gpyoung-at-dow.com


****************************************
Original message:

Hello Listers,

I wanted to see if anyone has had any experience scanning negatives with the Canon Canoscan 9900F. 
I have a low end Canoscan that I use for documents and the occasional photograph, which has given me very good results, but is not set up for negatives.  I have seen the recent posts about the Epson scanner, but I am trying to keep it under a $1000 so that I do not have to treat it as a capital purchase.  I have tried for the last two years to get a Coolscan 8000 on the capital budget with no luck, and do not see any scanner above $1000 as an option right now.

Thanks,
Steve


Steven Lee
Chief Technologists
Electron Microscopy Laboratory
Baylor University Medical Center
3500 Gaston Ave
Dallas, TX 750246
Ph: 214.820.3302
Fx: 214.820.4110
Em: stevenle-at-baylorhealth.edu



From MicroscopyL-request-at-ns.microscopy.com Wed May 19 02:44:19 2004



From: =?iso-8859-2?Q?KJ_H=FCbner?= :      hubner-at-IOd.krakow.pl
Date: Wed, 19 May 2004 08:50:23 +0200
Subject: [Microscopy] 2nd Iberian Conference on Pattern Recognition and Image Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


}
} IbPRIA'2005
}
} 2nd Iberian Conference on Pattern Recognition and Image Analysis
}
} June 7-9, 2005 - Estoril, Portugal
}
} http://ibpria2005.isr.ist.utl.pt
}
}
} =================
} ABOUT IbPRIA'2005
} =================
} IbPRIA'2005 is the second edition of the Iberian Conference on Pattern
} Recognition and Image Analysis and will be held in Estoril (Portugal)
} in June 7-9, 2005.
}
} IbPRIA is an international event co-organised every two years, by the
} Spanish and Portuguese Associations for Pattern Recognition (AERFAI
} and APRP). Its mission is to bring together researchers from all over
} the world working in Pattern Recognition and in all areas of Video,
} Image and Signal Processing. Participants will have the opportunity to
} meet colleagues working in these areas, to attend oral sessions and to
} exchange ideas during poster sessions.
}
} IbPRIA'2005 is sponsored by IAPR-International Association for Pattern
} Recognition and the conference proceedings are planned to be published
} by Springer in LNCS series.
}
} ======
} TOPICS
} ======
} The topics of the conference include, but are not limited to:
}
} - Pattern Recognition
} - Image Analysis
} - Computer Vision
} - Multimedia Systems
} - Statistical & Structural Pattern Recognition
} - Neural Networks
} - Bioinformatics
} - Image Coding and Processing
} - Shape and Texture Analysis
} - Biomedical Pattern Analysis
} - Information Systems
} - Speech Recognition
} - Natural Language Analysis
} - Document Processing
} - Robotics
} - Remote Sensing
} - Industrial Applications of Pattern Recognition
} - Special Hardware Architectures
}
} =================
} PROGRAM COMMITTEE
} =================
} Jake Aggarwal, University of Texas, USA
} José Benedi, Polyt. University of Valencia, Spain
} Isabelle Bloch, ENST, France
} Hervé Bourlard, EPFL, Switzerland
} Patrick Bouthemy, IRISA, France
} Horst Bunke, University of Bern, Switzerland
} Aurélio Campilho, University of Porto, Portugal
} Gilles Celeux, Université Paris-Sud, France
} Luigi Cordella, University of Napoles, Italy
} Alberto Del Bimbo, University of Firenze, Italy
} Hervé Delinguette, INRIA, France
} Rachid Deriche, INRIA, France
} José Dias, Instituto Superior Técnico, Portugal
} Robert Duin, University of Delft, The Netherlands
} Mário Figueiredo, Instituto Superior Técnico, Portugal
} Ana Fred, Instituto Superior Técnico, Portugal
} Andrew Gee, University of Cambridge, UK
} Mohamed Kamel, University of Waterloo, Canada
} Aggelos Katsaggelos, Northwestern University, USA
} Joseph Kittler, University of Surrey, UK
} Seong-Whan Lee, University of Korea, Korea
} Ana Mendonça, University of Porto, Portugal
} Hermann Ney, University of Aachen, Germany
} Wiro Niessen, University of Utrecht, The Netherlands
} Maria Petrou, University of Surrey, UK
} Armando Pinho, University of Aveiro, Portugal
} Ioannis Pitas, University of Tessaloniki, Greece
} Filiberto Pla, University Jaume I, Spain
} Richard Prager, University of Cambridge, UK
} José Principe, University of Florida, USA
} Ian Reid, University of Oxford, UK
} Gabriella Sanniti di Baja, Istituto di Cibernética, Italy
} Beatriz Santos, University of Aveiro, Portugal
} Jose Santos-Victor, Inst. Superior Técnico, Portugal
} Joan Serrat, Aut. University of Barcelona, Spain
} Yoshiaki Shirai, Osaka University, Japan
} Pierre Soille, Joint Research Centre, Italy
} Karl Tombre, LORIA, France
} M. Ines Torres, Univ. of the Basque Country, Spain
} Emmanuele Trucco, Heriot-Watt University, UK
} Alessandro Verri, University of Genova, Italy
} Max Viergever, University of Utrecht, The Netherlands
} Joachim Weickert, Saarland University, Germany
}
} ======
} CHAIRS
} ======
} GENERAL CO-CHAIRS:
} Jorge S. Marques (Instituto Superior Técnico, Portugal)
} Nicolás Pérez de la Blanca (University of Granada, Spain)
}
} LOCAL CHAIR:
} Pedro Pina (Instituto Superior Técnico, Portugal)
}
} ====================
} ORGANISING COMMITTEE
} ====================
} João Sanches, João Paulo Costeira, Teresa Barata, Vitorino Ramos,
} José Saraiva, Michele Mengucci, Fernando Soares
}
}
} ===================================
} PAPER SUBMISSION AND REVIEW PROCESS
} ===================================
} Papers should describe original and unpublished work on the topics of
} the conference. Prospective authors should prepare a full paper,
} written in english, not exceeding 8 pages and must submit it
} electronically.
}
} All manuscripts will be blind-reviewed by at least two members of the
} program committee. Accepted papers are intended to appear in the
} Springer LNCS series which will be distributed to all participants at
} the Conference.
}
} Submission implies that at least one of the authors has to register
} and to present the communication at the conference if the paper is
} accepted.
}
} ===============
} IMPORTANT DATES
} ===============
} Submission of papers : Noverber 12, 2004
} Notification of acceptance: January 7, 2005
} Camera-ready : January 30, 2005
} Early registration : April 7, 2005
} Conference : June 7-9, 2005
}
} =====
} VENUE
} =====
} The conference will take place in Estoril, a beautiful village located
} at about 20 km of Lisbon centre.
}
} The conference will be held at a four star hotel (Hotel Eden), that
} disposes of adequate facilities for workshops and conferences.
}
} The international airport of Lisbon is about 40 minutes from the
} venue. There is a direct bus-shuttle from/to the airport to the Hotel
} every hour. Alternatively, participants can also easily reach the
} venue by train, car or taxi.
}
} ========
} CONTACTS
} ========
} IbPRIA'2005 Secretariat
} CVRM / Geo-Systems Centre
} Instituto Superior Técnico
} Av. Rovisco Pais
} 1049-001 Lisboa
} PORTUGAL
}
} URL : http://ibpria2005.isr.ist.utl.pt
} e-mail: ibpria2005-at-isr.ist.utl.pt
} Tel : +351-218417438
} Fax : +351-218417442
}



From MicroscopyL-request-at-ns.microscopy.com Wed May 19 10:22:00 2004



From: RCHIOVETTI-at-aol.com
Date: Wed, 19 May 2004 11:31:13 EDT
Subject: [Microscopy] Orlando Workshop: Cryomicrotomy & Cryoultramicrotomy for Materials Sciences

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Fellow Listmembers,

Members in the Greater Orlando, Florida area are cordially invited to attend
a two day mini-workshop on "Cryomicrotomy and Cryoultramicrotomy for Materials
Sciences."

When:
Tuesday, June 8 and Wednesday, June 9, 2004, 9:00am-4:00pm

Where:
Room 101, Materials Characterization Facility
Advanced Materials Processing & Analysis Center
University of Central Florida
12443 Research Parkway
Orlando, FL 32826

What:
A two-day mini-workshop with presentations, demonstrations and hands-on
sessions for cryomicrotomy and cryoultramicrotomy, including toolmaking (wire loops
and hair probes), glass knife making, evaluation of glass knives, care and
cleaning of diamond knives and operation of the cryomicrotome and
cryoultramicrotome.

Background:
These techniques are of special interest to anyone working with polymers or
other materials which could benefit from sectioning or surface "polishing" at
low temperatures (no embedding required). The sections may be used for TEM,
optical microscopy, IR spectroscopy and FTIR. The polished surface of the bulk
material is suitable for AFM, SEM, X-ray microanalysis and elemental mapping.

Important Info:
There is no charge for this workshop.
Meals and refreshments will be served!
Attendance is open to everyone for the presentations and demonstrations on
the first day. However, attendance is limited for the cryosectioning hands-on
sessions on the second day.

Contacts:
To RSVP and to reserve a spot for the hands-on sessions, please contact
either Ms. Kim Megaw at Boeckeler Instruments, Inc. ( {kim-at-boeckeler.com} ,
800.552.2262) or Ms. Nancy Crouch at ICMAS, Inc. ( {nancy-at-icmas.com} , 865.984.8058)

Sponsors and Organizers:
University of Central Florida
Advanced Materials Processing & Analysis Center
Materials Characterization Facility
RMC Products Group, Boeckeler Instruments, Inc.

Hope to see you there!

Bob Chiovetti
Senior Product Specialist
Boeckeler Instruments, Inc.


From MicroscopyL-request-at-ns.microscopy.com Wed May 19 10:32:23 2004



From: Dusevich, Vladimir :      DusevichV-at-umkc.edu
Date: Wed, 19 May 2004 10:41:30 -0500
Subject: [Microscopy] RE: Re: RE: Re: re: ESEM comparisons

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

For ESEM in wet mode low voltage beam is not suitable,
even at 5 kV noise level is too high.

Vladimir

} I agree that resolution is more dependent on sample than
} microscope. What a microscope is capable of doing with an
} "ideal" sample often does not relate to the real world of
} biological samples which are anything but ideal!
}
} To get really good resolution you need to use higher kV
} values. These same values will result in the primary
} electrons penetrating too far into the low density biological
} sample resulting in substantial loss of signal in general and
} se1 & se2's in particular. Resulting image will be low
} resolution and often of much lesser quality than the same
} sample taken at a lower magnification.
}
} The problem of kV can be partially overcome by using a
} more coherent beam with greater beam density, i.e. Field
} Emission. FE allows one to work at lower kV values resulting
} in a se yield that is still adequate while minimizing sample
} damage. Resulting images are likely to be higher resolution
} (sharper with more detail) than comparable images at similar
} kVs from a tungsten instrument.
}
} However, there is a point where you go from useful
} magnification to "empty " magnification regardless of the gun
} type. This is sample dependent and must be determined
} accordingly. So it is unlikely that you will get any
} additional information from a bacteria sample at 100k than at
} say 50K. However you will most likely get a much sharper
} image at 20k with FE than you could with a W filament.
}
} Debby
}
}
}
} Debby Sherman, Manager Phone: 765-494-6666
} Life Science Microscopy Facility FAX: 765-494-5896
} Purdue University E-mail: dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907 http://www3.agriculture.purdue.edu/microscopy
}
}
} On 5/18/04 8:55 AM, "Dusevich, Vladimir" {DusevichV-at-umkc.edu} wrote:
}
} }
} }
} }
} ----------------------------------------------------------------------
} } --------
} } The Microscopy ListServer -- Sponsor: The Microscopy
} Society of America
} } To Subscribe/Unsubscribe --
} } http://www.msa.microscopy.com/MicroscopyListserver
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} --------------------------------------------------------------
} --------------
} --}
} -
} }
} } In my experience medium/high magnifications for wet specimens are
} } limited mostly by specimens themselves and not by a type of
} emitter.
} } While I can get a nice picture of gold on carbon at x100k,
} for most of
} } my "real life" specimens magnification is limited by x20k (or less).
} }
} } Vladimir
} }
} } } We look mostly at hydrated biological samples. We have a tungsten
} } } gun XL30 ESEM TMP. For looking at bacteria and cells a
} FEGESEM would
} } } be superior in terms of reduced damage and improved resolution. I
} } } find it hard to get medium/high resolution with wet biological
} } } samples. I feel FEGESEM would make this a lot easier.
} } }
} } } Dave
} } }
} } } On Mon, 17 May 2004 17:21:33 +1200 Bryony James
} } } {b.james-at-auckland.ac.nz} wrote:
} } }
} } } }
} } } }
} } } }
} } }
} ---------------------------------------------------------------------
} } } -
} } } } --------
} } } } The Microscopy ListServer -- Sponsor: The Microscopy
} } } Society of America
} } } } To Subscribe/Unsubscribe --
} } } http://www.msa.microscopy.com/MicroscopyListserver
} } } } On-Line Help
} } } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } }
} } } --------------------------------------------------------------
} } } -----------------
} } } }
} } } } Hello listers,
} } } }
} } } } We are in the process of justifying, to the University, a
} } } proposal to
} } } } purchase an ESEM. In prepapring the business case I am
} } } including the
} } } } ten year financials for both FEG and Tungtsen filament
} } } machines. Given
} } } } that we have to break even in that time, including covering our
} } } } depreciation, the Tungsten instrument is easier to justify
} } } in a purely
} } } } financial sense.
} } } }
} } } } From a scientific and operational viewpoint, though, I
} was hoping
} } } } some of you could provide opinions on the relative merits of
} } } Tungsten or FEG.
} } } } Our current conventional SEM is a Philips XL-30 S-FEG so we
} } } already know
} } } } of the costs associated with the emitter.
} } } }
} } } } The range of samples we will be required to put in our
} ESEM will be
} } } } anything you oculd possibly imagine from the Faculties of
} } } Engineering,
} } } } Science and Medicine. Samples will inlcude food materials
} } } (big dairy
} } } } industry in New Zealand), polymers, conducting polymers,
} mammalian
} } } } tissue, bacteria, fruit and veg, fish and chips, superconductors,
} } } } ceramic coatings and bees knees. We will definitely need a
} } } hot-stage.
} } } }
} } } } If you would like to contact me off list with your
} opinions please
} } } } feel free.
} } } }
} } } } Kind regards
} } } } Bryony James
} } } } Research Centre for Surface and Materials Science Univetrsity of
} } } } Auckland New Zealand
} } } }
} } } }
} } } }
} } } } This incoming email to UWE has been independently scanned
} } } for viruses
} } } } and any virus detected has been removed using McAfee anti-virus
} } } } software
} } } }
} } }
} } } ----------------------------------------
} } } Patton, David
} } } Email: David.Patton-at-uwe.ac.uk
} } } "University of the West of England"
} } }
} } }
} } }
} } } This email has been independently scanned for viruses and
} any virus
} } } detected has been removed using McAfee anti-virus software
} } }
} } }
} } }
} }
} }
} }
}
}
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Wed May 19 14:24:44 2004



From: RCHIOVETTI-at-aol.com
Date: Wed, 19 May 2004 15:33:52 EDT
Subject: [Microscopy] Madison, WI Workshop: Cryoultramicrotomy for Materials Sciences

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Fellow Listmembers,

Members in the Greater Madison, Wisconsin area are cordially invited to
attend a two day mini-workshop on "Cryoultramicrotomy for Materials Sciences."

When:
Wednesday, June 2 and Thursday, June 3, 2004, 9:00am-4:00pm

Where:
Room 374, Truax Building
Madison Area Technical College
3550 Anderson Street
Madison, WI 53704

(Gated lot near Truax Bldg. has audio to Parking Department. Mention this
Workshop for entrance to parking lot. Enter left side of building, take main
elevator to third floor)

What:
A two-day mini-workshop with presentations, demonstrations and hands-on
sessions for cryoultramicrotomy, including toolmaking (wire loops and hair probes),
glass knife making, evaluation of glass knives, care and cleaning of diamond
knives and operation of the cryomicrotome and cryoultramicrotome.

Background:
These techniques are of special interest to anyone working with polymers or
other materials which could benefit from sectioning or surface "polishing" at
low temperatures (no embedding required). The sections may be used for TEM,
optical microscopy, IR spectroscopy and FTIR. The polished surface of the bulk
material is suitable for AFM, SEM, X-ray microanalysis and elemental mapping.

Important Info:
There is no charge for this workshop.
Meals and refreshments will be served!
Attendance is open to everyone for the presentations and demonstrations on
the first day. However, attendance is limited for the cryosectioning hands-on
sessions on the second day.

Contacts:
To RSVP and to reserve a spot for the hands-on sessions, please contact
either Kim Megaw at Boeckeler Instruments, Inc. ( {kim-at-boeckeler.com} , 800.552.2262)
or Michael Kostrna at Madison Area Technical College
( {mkostrna-at-matcmadison.edu} , 608.246.6762).

Sponsors and Organizers:
Madison Area Technical College
(Electron Microscopy)
RMC Products Group, Boeckeler Instruments, Inc.
Doug D'Arcy, Midwest Regional Representative

See you in Madison!

Bob Chiovetti
Senior Product Specialist
Boeckeler Instruments, Inc.


From MicroscopyL-request-at-ns.microscopy.com Wed May 19 17:00:14 2004



From: Doug Baldwin :      dougbaldwin-at-mindspring.com
Date: Wed, 19 May 2004 15:07:16 -0700
Subject: [Microscopy] Hi-Res Microprocessor Photos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Fellow List Members,

I¹d appreciate your collective wisdom and experiences for a project I¹m
starting. I need to create high resolution (250 Mb minimum) photo files of
current microprocessor die faces (pentium 4 type) to show the circuitry and
traces with as much detail as possible. I¹d like to do this using my Nikon
Optiphot and a yet to be purchased motorized positionable stage with
automation software to include my Fuji S2 still camera (34 Mb sized files
per image) with image stitching as an added option. I have extensive
experience in Photoshop and would use that to stitch the images if the
automation software package doesn¹t include that. The stage travel only
needs to be about 3" (75mm) in the X and Y dimensions.

I¹ve been reading about some of the software programs available to help
accomplish this. I¹ve come across Image-Pro Plus with the Scope-Pro plugin
and the software program Metamorph in my Google searches. I don¹t need
extensive filtering or analysis functions in the software, just the ability
to create multiple, tileable images in a repeatable manner.

What economical software and hardware (such as the MultiScan-4 Low Cost
Scanning Stage for example) do you have experience with in similar projects?

Do any of you have an older version of Image Pro Plus or similar software
program that is sitting abandoned and alone on a shelf that needs a new
home? How about a used basic scanning stage that¹s collecting dust?

Any and all advice would be appreciated. You can email me off-list if that¹s
easier.

Thanks,
Doug Baldwin
Baldwin Hi-Tech Photography
dougbaldwin-at-mindspring.com




From MicroscopyL-request-at-ns.microscopy.com Wed May 19 19:03:05 2004



From: Gary Gaugler :      gary-at-gaugler.com
Date: Wed, 19 May 2004 17:11:52 -0700
Subject: [Microscopy] Re: Hi-Res Microprocessor Photos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The P4 BGA devices are like PPC G3, G4 and G5
products and are flip chip BGAs. Between the
inverted die and the multi layer ceramic package
is a thin layer of silicone. Through this layer
are many (700+) solder vias between the die and
package. Once the silicone is removed (and assuming
that you get the die off), the top layer greatly
masks the underlying 5-8 layers (depending on part
timeframe of manufacture) since it is used for interface
to the ceramic package.

This top-most layer must be removed to begin to
get at the actual interconnect layers. Once removed,
the underlying layers can be exposed by plasma
depassivation. Or, you can try wet etching.

These types of processors are face down rather than
face up. This is because they are too hot compared
to earlier CPUs.

The other factor is that you do not need high rez
image segments to make a high rez final image. To get
any reasonable image detail, you will likely need to
shoot at 500X or 1000X. At these mags, your field of
view is somewhere around 300u (guessing). For a 0.5cm
die, you will likely take 1,000 images (guessing again).
You can do the math. The point is that for a final high
rez image, it is made up of many low rez image segments.
The other problem you will face at high mag is the problem
of low DOF. The solution is to shoot with a SEM at high
tilt angle and dynamic focus. Then colorize the image.

If you need motorized and programmed stage movement,
Prior E103 is a good candidate and Soft Imaging analySIS
Opti is a good driving software product. Neither of
these items are free.

Anyway, I have found that it is not worth the effort.
But of course, YMMV.

gary g.


At 03:07 PM 5/19/2004, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Thu May 20 04:11:37 2004



From: Alan Stone :      as-at-astonmet.com
Date: Thu, 20 May 2004 07:30:41 -0500
Subject: [Microscopy] Thanks to All RE: Legal/Ethics Safety Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Doug

given the complexity of what you are doing, would it not be possible to try another method. If the die faces are flat and of the right size, why not just use a very high resolution flat bed scanner and do the whole thing in one go.

I'm sure that HP, Epson, Canon all do better than 3000dpi (optical resolution. It might certainly be worth investigating if camera photography involves a horrendous amount of image stitching (especially given the edge distortion in even the best camera). The only problems may be the reflective light quality of the scanner or how much depth of field there would be in a scanner image if the die has lots of depth to its layers.

My apologies if I've missed some technical point.

Malc

Malcolm Haswell
e.m. unit
University of Sunderland
UK
e-mail: malcolm.haswell-at-sunderland.ac.uk




----- Original Message -----
} From: Doug Baldwin {dougbaldwin-at-mindspring.com}


My thanks to everyone who replied to my legal/ethics query. After my
client became a bit more educated regarding the health risk we discovered
and checked with their counsel, they will do the right thing. The twist
here was that they were analyzing competitors' products which are made
overseas, so there are issues of diplomacy.

Alan Stone




Alan Stone
ASTON Metallurgical Services Co., Inc.
200 Larkin Drive Ste A
Wheeling, IL 60090
847/353-8100
www.astonmet.com



From MicroscopyL-request-at-ns.microscopy.com Thu May 20 09:52:22 2004



From: Ron Doole :      ron.doole-at-materials.oxford.ac.uk
Date: Thu, 20 May 2004 16:02:22 +0100 (GMT Daylight Time)
Subject: [Microscopy] Logging instrument use

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi All,

I wish to set up a logging system to record instrument use.
Ideally it would log date and time users start and
finish work on a particular project on each of our
instruments so that we can assess use of our facility.

At present users fill in a log book when they use an
instrument. It is easy to get general data (25 sessions to
the page gives an idea of instrument use but not user). I
don't want to waste time transcribing the data from the 20
instrument log books into a computer. Neither do I want to
spend a lot of money on each machine to gather the
information, however, most machines already have a PC
attached running the instrument or some ancillary equipment
and these are all networked.

What systems are being used at other sites? Any suggestions
of inexpensive systems that would work on two sites 6 miles
apart?

Regards,
Ron

----------------------
Mr. R.C. Doole
Department of Materials,
University of Oxford.
Parks Road, Oxford. OX1 3PH. UK.
Phone +44 (0) 1865 273701
Fax +44 (0) 1865 283333
ron.doole-at-materials.ox.ac.uk
http://www-em.materials.ox.ac.uk/
*********************************




From MicroscopyL-request-at-ns.microscopy.com Thu May 20 15:49:26 2004



From: Elaine Humphrey :      ech-at-interchange.ubc.ca
Date: Thu, 20 May 2004 13:58:37 -0700
Subject: [Microscopy] sputter coater glass

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Everyone
We have a Nonotech Semprep2 sputter coater which is so old the
company no longer exists. The glass cylinder has now got one more
gouge in it and it will have to lose at least 3 cm to grind it out.
Does anyone have any experience with where to buy a new glass
cylinder?

Many thanks
Elaine

--
Dr. Elaine Humphrey
Director, BioImaging Facility
President, Microscopy Society of Canada
University of British Columbia
6270 University Blvd, mail-stop Botany
Vancouver, BC
CANADA, V6T 1Z4
Phone: 604-822-3354
FAX: 604-822-6089
e-mail: ech-at-interchange.ubc.ca
website: www.emlab.ubc.ca


From MicroscopyL-request-at-ns.microscopy.com Thu May 20 17:26:04 2004



From: mocherla-at-eng.fsu.edu (by way of MicroscopyListserver)
Date: Thu, 20 May 2004 17:35:32 -0500
Subject: [Microscopy] viaWWW: staining and grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (mocherla-at-eng.fsu.edu) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Thursday, May 20, 2004 at 10:43:53
---------------------------------------------------------------------------

Email: mocherla-at-eng.fsu.edu
Name: supriya

Organization: FSU

Title-Subject: [Microscopy] [Filtered] Liposome Grids

Question: Dear authors,

I really appreciate this group for their support in my work.

I observed that for negative staining, carbon coated Formvar grid was used mostly. I would like to know if there is any significance in using this particlular grid or Can I use Carbon support film on Copper or Gold grid.

Thanks!!

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu May 20 19:24:44 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Thu, 20 May 2004 17:39:02 -0700
Subject: [Microscopy] Re: viaWWW: staining and grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 20, 2004, at 3:35 PM, by way of MicroscopyListserver wrote:

} Title-Subject: [Microscopy] [Filtered] Liposome Grids
}
} Question: Dear authors,
}
} I really appreciate this group for their support in my work.
}
} I observed that for negative staining, carbon coated Formvar grid was
} used mostly. I would like to know if there is any significance in
} using this particlular grid or Can I use Carbon support film on Copper
} or Gold grid.
}
Dear Supriya,
The grids most often used for negative staining are copper grids on
which there is a layer of formvar and usually an additional layer of
carbon. These are often referred to as carbon-formvar grids or
carbon-coated formvar grids. It is much easier to use a carbon-formvar
support film than to use just a carbon film for support. The resulting
film is much stronger than plain carbon, and does not degrade the
images of negatively stained preps. The web sites of several of the EM
supply companies describe their carbon-formvar grids.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Thu May 20 20:04:13 2004



From: Doug Baldwin :      dougbaldwin-at-mindspring.com
Date: Thu, 20 May 2004 18:10:14 -0700
Subject: [Microscopy] Re: Hi-Res Microprocessor Photos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Malcolm,

Interestingly, I had the same thought and tried this technique. I stopped by
one of the area's premier graphics arts service bureaus with an 8000+ dpi
flat bed scanner. We put the open faced CPU chip on the glass and proceeded
to scan it. It was awful. The main problem was that the light source on the
scanner was off-axis to the scanning array. This renders the face of the
chip dark and worthless.

Photos of microprocessors benefit from coaxial lighting since the metal
parts and traces reflect the light straight back to the lens and light
source. Lighting that's even slightly off-axis shows a marked drop off in
detail and reflectivity of the die faces.

The high dpi flat bed scanners are set up for flat artwork such as paintings
and drawings where side lighting is fine for rendering all the detail but
not for reflective metal parts.

Doug


}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
----------------------------------------------------------------------------
--} -
}
} Doug
}
} given the complexity of what you are doing, would it not be possible to try
} another method. If the die faces are flat and of the right size, why not just
} use a very high resolution flat bed scanner and do the whole thing in one go.
}
} I'm sure that HP, Epson, Canon all do better than 3000dpi (optical resolution.
} It might certainly be worth investigating if camera photography involves a
} horrendous amount of image stitching (especially given the edge distortion in
} even the best camera). The only problems may be the reflective light quality
} of the scanner or how much depth of field there would be in a scanner image if
} the die has lots of depth to its layers.
}
} My apologies if I've missed some technical point.
}
} Malc
}
} Malcolm Haswell
} e.m. unit
} University of Sunderland
} UK
} e-mail: malcolm.haswell-at-sunderland.ac.uk
}
}
}
}
} ----- Original Message -----
} } From: Doug Baldwin {dougbaldwin-at-mindspring.com}
} Date: Wednesday, May 19, 2004 11:07 pm
} Subject: [Microscopy] Hi-Res Microprocessor Photos
}
} }
} }
} } -------------------------------------------------------------------
} } -----------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } AmericaTo Subscribe/Unsubscribe --
} } http://www.msa.microscopy.com/MicroscopyListserverOn-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html--------
} } -------------------------------------------------------------------
} } ----
} }
} } Hello Fellow List Members,
} }
} } Id appreciate your collective wisdom and experiences for a
} } project Im
} } starting. I need to create high resolution (250 Mb minimum) photo
} } files of
} } current microprocessor die faces (pentium 4 type) to show the
} } circuitry and
} } traces with as much detail as possible. Id like to do this using
} } my Nikon
} } Optiphot and a yet to be purchased motorized positionable stage with
} } automation software to include my Fuji S2 still camera (34 Mb
} } sized files
} } per image) with image stitching as an added option. I have extensive
} } experience in Photoshop and would use that to stitch the images if the
} } automation software package doesnt include that. The stage travel
} } onlyneeds to be about 3" (75mm) in the X and Y dimensions.
} }
} } Ive been reading about some of the software programs available to
} } helpaccomplish this. Ive come across Image-Pro Plus with the
} } Scope-Pro plugin
} } and the software program Metamorph in my Google searches. I dont need
} } extensive filtering or analysis functions in the software, just
} } the ability
} } to create multiple, tileable images in a repeatable manner.
} }
} } What economical software and hardware (such as the MultiScan-4 Low
} } CostScanning Stage for example) do you have experience with in
} } similar projects?
} }
} } Do any of you have an older version of Image Pro Plus or similar
} } softwareprogram that is sitting abandoned and alone on a shelf
} } that needs a new
} } home? How about a used basic scanning stage thats collecting dust?
} }
} } Any and all advice would be appreciated. You can email me off-list
} } if thats
} } easier.
} }
} } Thanks,
} } Doug Baldwin
} } Baldwin Hi-Tech Photography
} } dougbaldwin-at-mindspring.com
} }
} }
} }
} }
}
}
}




From MicroscopyL-request-at-ns.microscopy.com Thu May 20 22:16:30 2004



From: Rosemary White :      Rosemary.White-at-csiro.au
Date: Fri, 21 May 2004 13:24:05 +1000
Subject: [Microscopy] Re: Re: Hi-Res Microprocessor Photos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

There is a photographer here in Oz who produces absolutely brilliant images
by scanning objects on a flat-bed scanner. They are very high resolution,
look as good as the images of leaves, flowers, insects (even very shiny
ones) we get using a good digital camera - Olympus DP70 or ProgRes or
similar, on a dissecting microscope. He's developed some ways of using the
scanner to get such great images. Name is Stuart Owen Fox, go here
http://www.smartstate.qld.gov.au/images/icase4g_2.jpg for an example.
cheers,
Rosemary


} From: Doug Baldwin {dougbaldwin-at-mindspring.com}
} Date: Thu, 20 May 2004 18:10:14 -0700
} To: {microscopy-at-msa.microscopy.com}
} Cc: Malcolm Haswell {malcolm.haswell-at-sunderland.ac.uk}
} Subject: [Microscopy] Re: Hi-Res Microprocessor Photos
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Malcolm,
}
} Interestingly, I had the same thought and tried this technique. I stopped by
} one of the area's premier graphics arts service bureaus with an 8000+ dpi
} flat bed scanner. We put the open faced CPU chip on the glass and proceeded
} to scan it. It was awful. The main problem was that the light source on the
} scanner was off-axis to the scanning array. This renders the face of the
} chip dark and worthless.
}
} Photos of microprocessors benefit from coaxial lighting since the metal
} parts and traces reflect the light straight back to the lens and light
} source. Lighting that's even slightly off-axis shows a marked drop off in
} detail and reflectivity of the die faces.
}
} The high dpi flat bed scanners are set up for flat artwork such as paintings
} and drawings where side lighting is fine for rendering all the detail but
} not for reflective metal parts.
}
} Doug
}
}
} }
} }
} }
-----------------------------------------------------------------------------} }
-
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} } http://www.msa.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} ----------------------------------------------------------------------------
} --} -
} }
} } Doug
} }
} } given the complexity of what you are doing, would it not be possible to try
} } another method. If the die faces are flat and of the right size, why not just
} } use a very high resolution flat bed scanner and do the whole thing in one go.
} }
} } I'm sure that HP, Epson, Canon all do better than 3000dpi (optical
} } resolution.
} } It might certainly be worth investigating if camera photography involves a
} } horrendous amount of image stitching (especially given the edge distortion in
} } even the best camera). The only problems may be the reflective light quality
} } of the scanner or how much depth of field there would be in a scanner image
} } if
} } the die has lots of depth to its layers.
} }
} } My apologies if I've missed some technical point.
} }
} } Malc
} }
} } Malcolm Haswell
} } e.m. unit
} } University of Sunderland
} } UK
} } e-mail: malcolm.haswell-at-sunderland.ac.uk
} }
} }
} }
} }
} } ----- Original Message -----
} } } From: Doug Baldwin {dougbaldwin-at-mindspring.com}
} } Date: Wednesday, May 19, 2004 11:07 pm
} } Subject: [Microscopy] Hi-Res Microprocessor Photos
} }
} } }
} } }
} } } -------------------------------------------------------------------
} } } -----------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } } AmericaTo Subscribe/Unsubscribe --
} } } http://www.msa.microscopy.com/MicroscopyListserverOn-Line Help
} } } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html--------
} } } -------------------------------------------------------------------
} } } ----
} } }
} } } Hello Fellow List Members,
} } }
} } } Id appreciate your collective wisdom and experiences for a
} } } project Im
} } } starting. I need to create high resolution (250 Mb minimum) photo
} } } files of
} } } current microprocessor die faces (pentium 4 type) to show the
} } } circuitry and
} } } traces with as much detail as possible. Id like to do this using
} } } my Nikon
} } } Optiphot and a yet to be purchased motorized positionable stage with
} } } automation software to include my Fuji S2 still camera (34 Mb
} } } sized files
} } } per image) with image stitching as an added option. I have extensive
} } } experience in Photoshop and would use that to stitch the images if the
} } } automation software package doesnt include that. The stage travel
} } } onlyneeds to be about 3" (75mm) in the X and Y dimensions.
} } }
} } } Ive been reading about some of the software programs available to
} } } helpaccomplish this. Ive come across Image-Pro Plus with the
} } } Scope-Pro plugin
} } } and the software program Metamorph in my Google searches. I dont need
} } } extensive filtering or analysis functions in the software, just
} } } the ability
} } } to create multiple, tileable images in a repeatable manner.
} } }
} } } What economical software and hardware (such as the MultiScan-4 Low
} } } CostScanning Stage for example) do you have experience with in
} } } similar projects?
} } }
} } } Do any of you have an older version of Image Pro Plus or similar
} } } softwareprogram that is sitting abandoned and alone on a shelf
} } } that needs a new
} } } home? How about a used basic scanning stage thats collecting dust?
} } }
} } } Any and all advice would be appreciated. You can email me off-list
} } } if thats
} } } easier.
} } }
} } } Thanks,
} } } Doug Baldwin
} } } Baldwin Hi-Tech Photography
} } } dougbaldwin-at-mindspring.com
} } }
} } }
} } }
} } }
} }
} }
} }
}
}
}
}




From MicroscopyL-request-at-ns.microscopy.com Thu May 20 23:52:06 2004



From: Allan Mitchell :      allan.mitchell-at-stonebow.otago.ac.nz
Date: Fri, 21 May 2004 17:00:39 +1200
Subject: [Microscopy] Identify Asbestos in the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all

Can anyone suggest a good text or reference for helping to identify
asbestos bodies in lung using an SEM, and using EDS.

Many thanks

Regards

Allan


From MicroscopyL-request-at-ns.microscopy.com Fri May 21 00:26:15 2004



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Fri, 21 May 2004 07:39:05 +0200
Subject: [Microscopy] sputter coater glass

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear Elaine

A good glassblower can sort it.
Either by producing another or taking the gouge out.
I have done that in the past on a VERY OLD Edwards. Definitely cheaper than the scientific company's. The quality of the work was good and the new dome is still in operation after 6 years.

-----Original Message-----
} From: Elaine Humphrey [mailto:ech-at-interchange.ubc.ca]
Sent: Thursday, May 20, 2004 10:59 PM
To: microscopy-at-msa.microscopy.com

Hello Everyone
We have a Nonotech Semprep2 sputter coater which is so old the
company no longer exists. The glass cylinder has now got one more
gouge in it and it will have to lose at least 3 cm to grind it out.
Does anyone have any experience with where to buy a new glass
cylinder?

Many thanks
Elaine

--
Dr. Elaine Humphrey
Director, BioImaging Facility
President, Microscopy Society of Canada
University of British Columbia
6270 University Blvd, mail-stop Botany
Vancouver, BC
CANADA, V6T 1Z4
Phone: 604-822-3354
FAX: 604-822-6089
e-mail: ech-at-interchange.ubc.ca
website: www.emlab.ubc.ca




From MicroscopyL-request-at-ns.microscopy.com Fri May 21 03:55:22 2004



From: Denan Konjhodzic :      denan-at-mpi-muelheim.mpg.de
Date: Fri, 21 May 2004 10:10:09 +0100
Subject: [Microscopy] AFM Phase images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Doug,

Regarding stitching images, we have successfully stitched micrographs using
PanaVue ImageAssembler (www.panavue.com) as it works with flat perspective.
Have not tried to stitch very large numbers of images though.

With regards
Thor


Dr Thor Bostrom
Analytical EM Facility; and
School of Physical and Chemical Sciences,
Queensland University of Technology (QUT)
Brisbane, QLD 4001, Australia

----------Original
message-------------------------------------------------------

Hi all!

does anybody know how is the phase in AFM software Nanoscope III for
Dimension 2100 defined? Could the bright regions in the phase image be
assigned to the depressions, if we are working in the attractive force
regime? Thank's a lot for any hint.

Best regards!

--
__________________________________________________

Ðenan Konjhodzic
Max-Planck-Institut für Kohlenforschung
Kaiser-Wilhelm-Platz 1
D-45470 Mülheim an der Ruhr

Phone: +49-208-306 2449
Fax: +49-208-306 2995
Mobile: +49-179-94 98 435
e-mail: denan-at-mpi-muelheim.mpg.de
___________________________________________________




From MicroscopyL-request-at-ns.microscopy.com Fri May 21 06:54:26 2004



From: Chaoying Ni :      cni-at-UDel.Edu
Date: Fri, 21 May 2004 08:02:43 -0400 (EDT)
Subject: [Microscopy] Re: Identify Asbestos in the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Allan,

I seriously doubt there is one as to unambiguously identify fibrous
asbestos is the job of TEM and (not or) an EDS on TEM.

****************************************
Chaoying Ni, PhD
The W.M. Keck Electron Microscope Facility
Facility location: 022 Spencer Laboratory
Mailing address:
201 duPont Hall
Dept of Matls Sci & Eng
University of Delaware
Newark, DE 19716
(302) 831-8354 (O); -2318(L); -4545(Fax)
http://eml.masc.udel.edu
*****************************************

On Fri, 21 May 2004, Allan Mitchell wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Hi all
}
} Can anyone suggest a good text or reference for helping to identify
} asbestos bodies in lung using an SEM, and using EDS.
}
} Many thanks
}
} Regards
}
} Allan
}


From MicroscopyL-request-at-ns.microscopy.com Fri May 21 07:57:16 2004



From: Larsen, Michael (Research) :      larsen-at-crd.ge.com
Date: Fri, 21 May 2004 09:06:14 -0400
Subject: [Microscopy] TEM of cells in solution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have some cells in solution. Is there some general technique to preserve
them so that I can put some down on a carbon membrane coated TEM grid
without them shriveling up? Can anyone refer me to a good reference on TEM
sample prep of cells?


From MicroscopyL-request-at-ns.microscopy.com Fri May 21 08:09:47 2004



From: Bryan :      bbandli-at-mvainc.com
Date: Fri, 21 May 2004 09:16:35 -0400
Subject: [Microscopy] Re: Identify Asbestos in the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Allan,

The Mineralogical Society of America has a series of books on
mineralogical and geochemical topics. Volume 28 is titled "Health
Effects of Mineral Dusts" and covers a wide range of topics, from basic
asbestos mineralogy to cellular and molecular mechanisms for disease.
I'm not sure if it specifically deals with identifying asbestos bodies
in lung tissue specificaly but it should have some references, and there
should be useful information to help with EDS identification. It's
available at the MSA (the other MSA) website at
www.minsocam.org/MSA/RIM/ for a very reasonable price for the amount of
information contained ($32 US).

Best Regards,
Bryan Bandli

Allan Mitchell wrote:

}
}
} ------------------------------------------------------------------------------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
}
} Hi all
}
} Can anyone suggest a good text or reference for helping to identify
} asbestos bodies in lung using an SEM, and using EDS.
}
} Many thanks
}
} Regards
}
} Allan
}
}

--
-----------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------
The information in this email is confidential and may be legally privileged.
It is intended solely for the addressee. If you are not the intended recipient, please delete the email and notify MVA Scientific Consultants of the transmission error.
MVA Scientific Consultants - 5500 Oakbrook Pkwy Suite 200, Norcross, GA 30093 - (770)662-8509
-------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------





From MicroscopyL-request-at-ns.microscopy.com Fri May 21 10:48:40 2004



From: Monson, Frederick :      fmonson-at-wcupa.edu
Date: Fri, 21 May 2004 11:56:02 -0400
Subject: [Microscopy] Identify Asbestos in the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I had a request recently to perform EDS on a presumed crysotile
crystal/shard that had been located with TEM on a grid. We have an
Oxford INCA system on our FEI Quanta 400, so I resurrected an old grid
holder from a salvaged RCA EMU4, mounted the grid in holder in the ESEM,
located the offending grid square and performed a point and ID on the
crystal in question.

As Chaoying suggests, this would not be in concert with NIOSH 7402
(TEM)
http://www.cdc.gov/niosh/nmam/pdfs/7402.pdf, or NIOSH 9000 (XRD)
http://www.cdc.gov/niosh/nmam/pdfs/7402.pdf, identifications of
asbestos, to give quick examples.

Other sources and possibilities for your problem(?):
SEM+:
http://sup.ultrakohl.com/News-nov/mesoth.htm

EBSD:

http://www.asbestostemlabs.com/staff.htm

SEM/EBSD
http://www.dxcicdd.com/01/pdf/D-075.pdf

I don't believe that EBSD/SEM has been approved for asbestos:

http://hyperphysics.phy-astr.gsu.edu/hbase/minerals/asbestos.html

But that doesn't mean that SEM can't be used in simple identifications
of morphology and composition. The real problem is to find it, if you
are looking in tissue. Mapping for the relevant elements should work,
but for very small crystals, you will be working at or above the 'good
result' window for EDS and mapping when you want to identify and image
at high mag.

Hope some of this helps,

Fred


Frederick C. Monson, PhD
Center for Advanced Scientific Imaging
Mail to Geology
West Chester University of Pennsylvania
Schmucker II Science Center, Room SS024
South Church Street and Rosedale Avenue
West Chester, PA, 19383
Phone/FAX: 610-738-0437
eMail: fmonson-at-wcupa.edu
CASI Page and Scheduling
http://darwin.wcupa.edu/CASI/


-----Original Message-----
} From: Allan Mitchell [mailto:allan.mitchell-at-stonebow.otago.ac.nz]
Sent: Friday, May 21, 2004 1:01 AM
To: microscopy-at-msa.microscopy.com

Hi all

Can anyone suggest a good text or reference for helping to identify
asbestos bodies in lung using an SEM, and using EDS.

Many thanks

Regards

Allan




From MicroscopyL-request-at-ns.microscopy.com Fri May 21 11:05:04 2004



From: Tamara Howard :      thoward-at-unm.edu
Date: Fri, 21 May 2004 10:13:49 -0600 (MDT)
Subject: [Microscopy] Re: TEM of cells in solution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Do you want to just look at them whole, in a HVEM? No
embedding/sectioning? What do you want to see?

On Fri, 21 May 2004, Larsen, Michael (Research) wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} I have some cells in solution. Is there some general technique to preserve
} them so that I can put some down on a carbon membrane coated TEM grid
} without them shriveling up? Can anyone refer me to a good reference on TEM
} sample prep of cells?
}
}

|--------------------------------------------------|
Tamara Howard
Department of Cell Biology and Physiology
University of New Mexico - Health Sciences Center
Albuquerque, NM 87131
thoward-at-unm.edu
|--------------------------------------------------|



From MicroscopyL-request-at-ns.microscopy.com Fri May 21 11:51:52 2004



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Fri, 21 May 2004 09:58:51 -0700
Subject: [Microscopy] Re: Re: Hi-Res Microprocessor Photos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Today's microprocessors have micron scale metal lines. You would require submicron resolution to distinguish the lines. A flat bed scanner that would resolve micron scale lines would need at least 25,000 DPI resolution just to detect the lines.

John Mardinly
Intel

-----Original Message-----
} From: Rosemary White [mailto:Rosemary.White-at-csiro.au]
Sent: Thursday, May 20, 2004 8:24 PM
To: Doug Baldwin; microscopy-at-msa.microscopy.com
Cc: Malcolm Haswell

There is a photographer here in Oz who produces absolutely brilliant images
by scanning objects on a flat-bed scanner. They are very high resolution,
look as good as the images of leaves, flowers, insects (even very shiny
ones) we get using a good digital camera - Olympus DP70 or ProgRes or
similar, on a dissecting microscope. He's developed some ways of using the
scanner to get such great images. Name is Stuart Owen Fox, go here
http://www.smartstate.qld.gov.au/images/icase4g_2.jpg for an example.
cheers,
Rosemary


} From: Doug Baldwin {dougbaldwin-at-mindspring.com}
} Date: Thu, 20 May 2004 18:10:14 -0700
} To: {microscopy-at-msa.microscopy.com}
} Cc: Malcolm Haswell {malcolm.haswell-at-sunderland.ac.uk}
} Subject: [Microscopy] Re: Hi-Res Microprocessor Photos
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Malcolm,
}
} Interestingly, I had the same thought and tried this technique. I stopped by
} one of the area's premier graphics arts service bureaus with an 8000+ dpi
} flat bed scanner. We put the open faced CPU chip on the glass and proceeded
} to scan it. It was awful. The main problem was that the light source on the
} scanner was off-axis to the scanning array. This renders the face of the
} chip dark and worthless.
}
} Photos of microprocessors benefit from coaxial lighting since the metal
} parts and traces reflect the light straight back to the lens and light
} source. Lighting that's even slightly off-axis shows a marked drop off in
} detail and reflectivity of the die faces.
}
} The high dpi flat bed scanners are set up for flat artwork such as paintings
} and drawings where side lighting is fine for rendering all the detail but
} not for reflective metal parts.
}
} Doug
}
}
} }
} }
} }
-----------------------------------------------------------------------------} }
-
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} } http://www.msa.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} }
} ----------------------------------------------------------------------------
} --} -
} }
} } Doug
} }
} } given the complexity of what you are doing, would it not be possible to try
} } another method. If the die faces are flat and of the right size, why not just
} } use a very high resolution flat bed scanner and do the whole thing in one go.
} }
} } I'm sure that HP, Epson, Canon all do better than 3000dpi (optical
} } resolution.
} } It might certainly be worth investigating if camera photography involves a
} } horrendous amount of image stitching (especially given the edge distortion in
} } even the best camera). The only problems may be the reflective light quality
} } of the scanner or how much depth of field there would be in a scanner image
} } if
} } the die has lots of depth to its layers.
} }
} } My apologies if I've missed some technical point.
} }
} } Malc
} }
} } Malcolm Haswell
} } e.m. unit
} } University of Sunderland
} } UK
} } e-mail: malcolm.haswell-at-sunderland.ac.uk
} }
} }
} }
} }
} } ----- Original Message -----
} } } From: Doug Baldwin {dougbaldwin-at-mindspring.com}
} } Date: Wednesday, May 19, 2004 11:07 pm
} } Subject: [Microscopy] Hi-Res Microprocessor Photos
} }
} } }
} } }
} } } -------------------------------------------------------------------
} } } -----------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } } AmericaTo Subscribe/Unsubscribe --
} } } http://www.msa.microscopy.com/MicroscopyListserverOn-Line Help
} } } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html--------
} } } -------------------------------------------------------------------
} } } ----
} } }
} } } Hello Fellow List Members,
} } }
} } } Id appreciate your collective wisdom and experiences for a
} } } project Im
} } } starting. I need to create high resolution (250 Mb minimum) photo
} } } files of
} } } current microprocessor die faces (pentium 4 type) to show the
} } } circuitry and
} } } traces with as much detail as possible. Id like to do this using
} } } my Nikon
} } } Optiphot and a yet to be purchased motorized positionable stage with
} } } automation software to include my Fuji S2 still camera (34 Mb
} } } sized files
} } } per image) with image stitching as an added option. I have extensive
} } } experience in Photoshop and would use that to stitch the images if the
} } } automation software package doesnt include that. The stage travel
} } } onlyneeds to be about 3" (75mm) in the X and Y dimensions.
} } }
} } } Ive been reading about some of the software programs available to
} } } helpaccomplish this. Ive come across Image-Pro Plus with the
} } } Scope-Pro plugin
} } } and the software program Metamorph in my Google searches. I dont need
} } } extensive filtering or analysis functions in the software, just
} } } the ability
} } } to create multiple, tileable images in a repeatable manner.
} } }
} } } What economical software and hardware (such as the MultiScan-4 Low
} } } CostScanning Stage for example) do you have experience with in
} } } similar projects?
} } }
} } } Do any of you have an older version of Image Pro Plus or similar
} } } softwareprogram that is sitting abandoned and alone on a shelf
} } } that needs a new
} } } home? How about a used basic scanning stage thats collecting dust?
} } }
} } } Any and all advice would be appreciated. You can email me off-list
} } } if thats
} } } easier.
} } }
} } } Thanks,
} } } Doug Baldwin
} } } Baldwin Hi-Tech Photography
} } } dougbaldwin-at-mindspring.com
} } }
} } }
} } }
} } }
} }
} }
} }
}
}
}
}








From MicroscopyL-request-at-ns.microscopy.com Fri May 21 14:17:27 2004



From: Philip Oshel :      peoshel-at-wisc.edu
Date: Fri, 21 May 2004 14:26:31 -0500
Subject: [Microscopy] Gatan cryoholder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Listers,

I'll try again. Does anyone have a Gatan model 626-300 cryoholder to
sell or donate to a university?
Thanks.

Phil
--
Philip Oshel
Supervisor, BBPIC microscopy facility
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706 - 1284
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)


From MicroscopyL-request-at-ns.microscopy.com Fri May 21 14:21:00 2004



From: Marek Malecki, M.D., Ph.D., Professor :      mmalecki-at-wisc.edu
Date: Fri, 21 May 2004 14:28:45 -0500
Subject: [Microscopy] wanted: epi-fluorescence lamp housing and slider for Diaphot

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We would like to buy or trade in:
epi-fluorescence lamp housing and slider for Diaphot (inverted Nikon)
new or refurbished.

Vendors welcome!



From MicroscopyL-request-at-ns.microscopy.com Fri May 21 14:48:08 2004



From: Klughammer :      schmaus-at-klughammer.de
Date: Fri, 21 May 2004 21:55:18 +0200
Subject: [Microscopy] Re: Hi-Res Microprocessor Photos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Doug,

Image-Pro Plus is a very good image analysis software with
extensive capabilities. But there is a another product from
Media Cybernetics which is "Image-Pro Discovery". This
is the smaller version of Image-Pro Plus and therefore
cheaper. It allows you to capture X-Y automatically in
combination with Scope-Pro. The software gives you the
possibility to select the order how you want to capture
the images and it stitches them automatically.
You also could capture Z-stacks automatically either by
moving the microscope stage or with a motorized objective
(Piezo) which gives you very precise results.

However if you want to work with an automatic x-y stage you
should work with a camera which is integrated in the
capturing software because then the software also captures
the images, then the stage moves on, the image is captured and so
on. With your Fuji camera you have to do the capturing
manually which is very cumbersome. You should use a digital
CCD camera which is connected to your computer and shows you
the live image on the computer monitor.

Please let me know if you have further questions.

mfg / regards

Anneliese Schmaus
Product Manager

klughammer gmbh
Strassbach 9
85229 Markt Indersdorf
Germany
Tel. +49 08136 6011
Fax +49 08136 7098
info-at-klughammer.de
www.klughammer.de



DB} ------------------------------------------------------------------------------
DB} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
DB} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
DB} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
DB} -------------------------------------------------------------------------------

DB} Hello Fellow List Members,

DB} I¹d appreciate your collective wisdom and experiences for a project I¹m
DB} starting. I need to create high resolution (250 Mb minimum) photo files of
DB} current microprocessor die faces (pentium 4 type) to show the circuitry and
DB} traces with as much detail as possible. I¹d like to do this using my Nikon
DB} Optiphot and a yet to be purchased motorized positionable stage with
DB} automation software to include my Fuji S2 still camera (34 Mb sized files
DB} per image) with image stitching as an added option. I have extensive
DB} experience in Photoshop and would use that to stitch the images if the
DB} automation software package doesn¹t include that. The stage travel only
DB} needs to be about 3" (75mm) in the X and Y dimensions.

DB} I¹ve been reading about some of the software programs available to help
DB} accomplish this. I¹ve come across Image-Pro Plus with the Scope-Pro plugin
DB} and the software program Metamorph in my Google searches. I don¹t need
DB} extensive filtering or analysis functions in the software, just the ability
DB} to create multiple, tileable images in a repeatable manner.

DB} What economical software and hardware (such as the MultiScan-4 Low Cost
DB} Scanning Stage for example) do you have experience with in similar projects?

DB} Do any of you have an older version of Image Pro Plus or similar software
DB} program that is sitting abandoned and alone on a shelf that needs a new
DB} home? How about a used basic scanning stage that¹s collecting dust?

DB} Any and all advice would be appreciated. You can email me off-list if that¹s
DB} easier.

DB} Thanks,
DB} Doug Baldwin
DB} Baldwin Hi-Tech Photography
DB} dougbaldwin-at-mindspring.com



From MicroscopyL-request-at-ns.microscopy.com Fri May 21 15:19:40 2004



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Fri, 21 May 2004 13:27:20 -0700
Subject: [Microscopy] Re: Re: viaWWW: staining and grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Bill
I disagree with you: carbon coated Formvar film do degrade the image in
your words versus just carbon. Formvar film is quite thick in compare with
"plain carbon", so it "adsorbs" more electrons and therefore "degrade the
image"... Another problem with Formvar and other plastic support films is
that they change their properties under the beam which cause the image
drift... So, in my point of view, it's just wrong statement that carbon
coated Formvar film "does not degrade the images of negatively stained
preps". It's quite opposite: pure thin carbon film is very stable under
the beam, much thinner than any plastic film and therefore provides best
possible (for biological samples) support material in terms of stability
and "transparency" (to the electrons) for negative staining. Carbon coated
plastic support films are popular just because it's much easier to produce
in the Lab (and not necessary the best choice). In my Lab we routinely use
1.4-1.8 nm thick "pure" carbon support films over "holey" film. You may
not use such thin film for huge objects like whole cell, but it works great
for macromolecules of any sort. "Carbon over holey film" grids are
available thorough major EM suppliers. Interestingly, I never used carbon
coated Formvar films. I used to use carbon coating on cellulose-derivative
films like Parlodion (because it drifted so much without carbon) and I used
"naked" Formvar film because it's more stable under the beam than other
variety of films. Formvar film looks much "darker" in the scope than
Parlodion, so I prefer to use carbon coated Parlodion film if I could not
use "normal carbon": very contaminated samples with a lot of aggregates,
cells etc. Someone may need to try "naked" Parlodion (or similar) because
good hydrophilic properties of cellulose-deviated films if Formvar of
carbon is too hydrophobic to them (at the huge price of film instability in
the beam). I don't like glow-discharge, personally: I would rather use
poly-lysine coating. Have a great weekend, Sergey


At 05:39 PM 5/20/2004 -0700, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Fri May 21 16:18:31 2004



From: David Elliott :      David.Elliott-at-yale.edu
Date: Fri, 21 May 2004 17:27:43 -0400
Subject: [Microscopy] Sensys camera value

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi listers
I have a problem I am hoping that someone out there can help me with.
7 years ago our lab jointly bought a Photometrix SenSys PVCAM with
another lab. The labs are splitting up. We need to determine the
current value of the camera so that one lab can pay the other lab. Can
anybody point me to this data?
Thanx
David



From MicroscopyL-request-at-ns.microscopy.com Fri May 21 17:29:47 2004



From: Thomas H. Kosel :      kosel-at-nd.edu
Date: Fri, 21 May 2004 17:38:54 -0500
Subject: [Microscopy] RE: TEM Film scanner Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi -

We have been using the Epson Perfection 3200 scanner for 8x10 cm TEM
negatives for about a year now, and find it excellent. The
resolution (3200 dpi optical) is almost as good as the 4000 dpi on
the Nikon 8000 etc., and we successfully scanned some very dense
(accidentally overexposed) negatives when our TEM's exposure meter
went on the fritz. We mount the negatives in the standard 4x5"
negative carrier. (The Nikon negative holder doesn't fit 8x10 cm
negatives without modification, or cutting the negatives down to
size.) The 3200 was listed at $299 on the Epson website the last
time I looked.

Cheers,
Tom Kosel


Original message:

Hello Listers,

I wanted to see if anyone has had any experience scanning negatives
with the Canon Canoscan 9900F.
I have a low end Canoscan that I use for documents and the occasional
photograph, which has given me
very good results, but is not set up for negatives. I have seen the
recent posts about the Epson scanner,
but I am trying to keep it under a $1000 so that I do not have to
treat it as a capital purchase. I have tried
for the last two years to get a Coolscan 8000 on the capital budget
with no luck, and do not see any scanner
above $1000 as an option right now.

Thanks,
Steve


Steven Lee
Chief Technologists
Electron Microscopy Laboratory
Baylor University Medical Center
3500 Gaston Ave
Dallas, TX 750246
Ph: 214.820.3302
Fx: 214.820.4110
Em: stevenle-at-baylorhealth.edu
--
*********************************
Thomas H. Kosel
Department of Electrical Engineering
University of Notre Dame
Notre Dame, IN 46556
(574) 631-5642 (201 Cushing)
(574) 631-4393 FAX
kosel-at-nd.edu
*********************************


From MicroscopyL-request-at-ns.microscopy.com Fri May 21 19:08:23 2004



From: jbai :      jbai-at-ic.sunysb.edu
Date: Sat, 22 May 2004 00:17:20 GMT
Subject: [Microscopy] Reference on the determination of dislocation line direction by TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,


I want to find some papers or books which introduce the determination of
the direction of dislocation lines by TEM.

Thank you very much,

Regards,

Nina



From MicroscopyL-request-at-ns.microscopy.com Sat May 22 02:05:24 2004



From: Larry Stoter :      larry-at-cymru.freewire.co.uk
Date: Sat, 22 May 2004 08:02:13 +0100
Subject: [Microscopy] Re: Reference on the determination of dislocation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

"Electron Microscopy of Thin Crystals"
Hirsch, Howie, Nicholson, Pashley & Whelan
Pub: Robert E. Krieger
ISBN: 0-88275-376-2


"Transmission Electron Microscopy of Materials"
Thomas & Goringe
Pub: John Wiley & Sons
ISBN: 0-471-12244-0

"Defect Analysis in Electron Microscopy"
Loretto & Smallman
Pub: Chapman & Hall
ISBN: 0-470-54760-X

The latter, in particular, is quite detailed about dislocation
analysis with specific instructions. It also gives details of
weak-beam methods, which give much higher resolution images of
dislocations, very useful if you have closely packed dislocations.

Unfortunately, I suspect all are no longer in print.
--
Larry Stoter
PLEASE NOTE
1. Any mail other than plain text will be automatically deleted.
2. Any mail, legitimate or not, apparently or actually from hotmail
will automatically be deleted.
:-)


From MicroscopyL-request-at-ns.microscopy.com Sat May 22 04:58:19 2004



From: Juha Dapper :      juha_dapper-at-yahoo.com
Date: Sat, 22 May 2004 03:07:31 -0700 (PDT)
Subject: [Microscopy] HRTEM: how resolutions defined in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear microscopist experts,

When I read literatures, I have been all the time
confused by the resolution definition in transmission
electron microscopy. Could you please kindly help?

For a FEG-HRTEM, what is the "line resolution" as
opposed to the "information resolution"? What is the
difference between each other?

How could I achieve the "line resolution" and the
"information resolution" in TEM experiments? And, how
can I resolve (identify, or "see") the "line
resolution" and the "information resolution"
respectively from a state-of-the-art HREM image?

Thank you for your time. Wish you a nice weekend,

-Juha





__________________________________
Do you Yahoo!?
Yahoo! Domains – Claim yours for only $14.70/year
http://smallbusiness.promotions.yahoo.com/offer


From MicroscopyL-request-at-ns.microscopy.com Sat May 22 09:13:06 2004



From: antuni-at-aol.com (by way of MicroscopyListserver)
Date: Sat, 22 May 2004 09:22:31 -0500
Subject: [Microscopy] viaWWW: Identifying Asbestos Using ptical Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (antuni-at-aol.com) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday, May 21, 2004 at 16:19:26
---------------------------------------------------------------------------

Email: antuni-at-aol.com
Name: Anthony Ribaudo

Organization: Consultant

Title-Subject: [Microscopy] [Filtered] Identifying Asbestos Using ptical Microscopy

Question: Can one identify asbestos only utilizing optical microscopy without the aid of electron diffraction/TEM ?


Anthony Ribaudo



---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sat May 22 09:14:50 2004



From: christian_b_a-at-hotmail.com (by way of Ask-A-Microscopist)
Date: Sat, 22 May 2004 09:24:15 -0500
Subject: [Microscopy] AskAMicroscopist: degrade the equipment lifetime

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (christian_b_a-at-hotmail.com) from http://www.msa.microscopy.org/Ask-A-Microscopist.html on Friday, May 21, 2004 at 09:07:25
---------------------------------------------------------------------------

Email: christian_b_a-at-hotmail.com
Name: christian albano

Organization: Neuropsychiatric Research Institute

Education: Graduate College

Location: Fargo, ND, USA

Question: Does using the Fluoview Olympus LSCM for non-floresence degrade the equipment lifetime?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sat May 22 11:52:37 2004



From: Chiphead :      chiphead-at-sbcglobal.net
Date: Sat, 22 May 2004 12:00:58 -0500
Subject: [Microscopy] viaWWW: Identifying Asbestos Using ptical

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Warning: Non scientific answer:

I remember an edition of "This Old House" or one of those types of shows
where they took some shingle samples to a lab for testing.

The lab used polarized microscopy to do at least an "initial" screening
and determine that the shingles "most likely" contained asbestos.

The "would do some additional testing" to verify (it was, and the house
needed to have a tent built around it, etc., etc.). I assume the
additional testing was something along the lines of TEM, etc.

The conclusion I drew was that if you had materials that you were
expecting asbestos, this (polarized examination) was a good screening
tool. If the fibers didn't change from one specific color to another as
the analyzer was rotated, you didn't have asbestos. If they did, you
probably have asbestos, but there are other fibers that may do the same
thing.

These were also fibers, not dust, and I don't know if that would affect
the results or not.

Sorry if this wasn't useful.

John R.

-----Original Message-----
} From: by way of MicroscopyListserver [mailto:antuni-at-aol.com]
Sent: Saturday, May 22, 2004 9:23 AM
To: microscopy-at-ns.microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (antuni-at-aol.com) from
http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday,
May 21, 2004 at 16:19:26
------------------------------------------------------------------------
---

Email: antuni-at-aol.com
Name: Anthony Ribaudo

Organization: Consultant

Title-Subject: [Microscopy] [Filtered] Identifying Asbestos Using ptical
Microscopy

Question: Can one identify asbestos only utilizing optical microscopy
without the aid of electron diffraction/TEM ?


Anthony Ribaudo



------------------------------------------------------------------------
---




From MicroscopyL-request-at-ns.microscopy.com Sat May 22 16:00:20 2004



From: Gordon Couger :      gcouger-at-provalue.net
Date: Sat, 22 May 2004 16:09:06 -0500
Subject: [Microscopy] Re: RE: viaWWW: Identifying Asbestos Using ptical Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Only one of the forms of asbestos is harmful to humans. Chrysotile
and then only if it is inhaled and the particle size is such it
lodges in the lung so it can form a site for cancer to get a start.
Chrysotile is only a small percentage of asbestos and the particle
size must fall in the 5 to 15 micron range to stay in the lungs.

These conditions are almost only found in asbestos worker and miners
and not in the general public. In the case of high exposure to
asbestos cigarette smoking is also a very large contributor to the
cancer caused by asbestos.

Had the asbestos studies been allowed to be fully peer reviewed
before legislation was passed we would probably still be using
asbestos in many applications today with more protection for
asbestos workers and minors and restrictions on some products such
as asbestos powder and asbestos cloth. Unfortunately the legislators
won't review thier laws once they are passed and the western world
is spending needless billions of dollars in asbestos abatement in
every old building that is renovated.

Gordon
Gordon Couger gcc-at-couger.com

I collect links on information related to light microscopes.
http://www.couger.com/microscope/links/gclinks.html
Please forward any links or information you think might be useful to
others.
Microscope Manual at www.science-info.org

} From: "Chiphead" {chiphead-at-sbcglobal.net}

:
: Warning: Non scientific answer:
:
: I remember an edition of "This Old House" or one of those types of
shows
: where they took some shingle samples to a lab for testing.
:
: The lab used polarized microscopy to do at least an "initial"
screening
: and determine that the shingles "most likely" contained asbestos.
:
: The "would do some additional testing" to verify (it was, and the
house
: needed to have a tent built around it, etc., etc.). I assume the
: additional testing was something along the lines of TEM, etc.
:
: The conclusion I drew was that if you had materials that you were
: expecting asbestos, this (polarized examination) was a good
screening
: tool. If the fibers didn't change from one specific color to
another as
: the analyzer was rotated, you didn't have asbestos. If they did,
you
: probably have asbestos, but there are other fibers that may do the
same
: thing.
:
: These were also fibers, not dust, and I don't know if that would
affect
: the results or not.
:
: Sorry if this wasn't useful.
:
: John R.
:
: -----Original Message-----
: } From: by way of MicroscopyListserver [mailto:antuni-at-aol.com]
: Sent: Saturday, May 22, 2004 9:23 AM
: To: microscopy-at-ns.microscopy.com
: Subject: [Microscopy] viaWWW: Identifying Asbestos Using ptical
: Microscopy
:
:
:
: ------------------------------------------------------------------
------
: ------
: The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
: To Subscribe/Unsubscribe --
: http://www.msa.microscopy.com/MicroscopyListserver
: On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
: ------------------------------------------------------------------
------
: -------
:
: Below is the result of your feedback form (NJZFM-ultra-55). It
was
: submitted by (antuni-at-aol.com) from
: http://microscopy.com/MicroscopyListserver/MLFormMail.html on
Friday,
: May 21, 2004 at 16:19:26
: ------------------------------------------------------------------
------
: ---
:
: Email: antuni-at-aol.com
: Name: Anthony Ribaudo
:
: Organization: Consultant
:
: Title-Subject: [Microscopy] [Filtered] Identifying Asbestos
Using ptical
: Microscopy
:
: Question: Can one identify asbestos only utilizing optical
microscopy
: without the aid of electron diffraction/TEM ?
:
:
: Anthony Ribaudo
:
:
:
: ------------------------------------------------------------------
------
: ---
:
:
:
:
:




From MicroscopyL-request-at-ns.microscopy.com Mon May 24 09:31:26 2004



From: Rosemary White :      Rosemary.White-at-csiro.au
Date: Mon, 24 May 2004 09:27:49 +1000
Subject: [Microscopy] asbestos risks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

This is off the topic of identifying asbestos, but thought I should comment
that just yesterday there was a program on the radio (streaming audio
available at http://www.abc.net.au/rn/talks/bbing/, transcript available by
Thursday) about new cases of asbestosis and mesothelioma showing up. The
young people (20s and 30s) now getting this lived in high asbestos areas as
kids, were children of builders (and builders themselves) who demolished or
renovated fibro houses, etc. Perhaps just an Australian problem....
Rosemary

} From: "Gordon Couger" {gcouger-at-provalue.net}
} Date: Sat, 22 May 2004 16:09:06 -0500
} To: "Chiphead" {chiphead-at-sbcglobal.net} , "'by way of MicroscopyListserver'"
} {antuni-at-aol.com} , {microscopy-at-ns.microscopy.com}
} Subject: [Microscopy] Re: RE: viaWWW: Identifying Asbestos Using ptical
} Microscopy
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Only one of the forms of asbestos is harmful to humans. Chrysotile
} and then only if it is inhaled and the particle size is such it
} lodges in the lung so it can form a site for cancer to get a start.
} Chrysotile is only a small percentage of asbestos and the particle
} size must fall in the 5 to 15 micron range to stay in the lungs.
}
} These conditions are almost only found in asbestos worker and miners
} and not in the general public. In the case of high exposure to
} asbestos cigarette smoking is also a very large contributor to the
} cancer caused by asbestos.
}
} Had the asbestos studies been allowed to be fully peer reviewed
} before legislation was passed we would probably still be using
} asbestos in many applications today with more protection for
} asbestos workers and minors and restrictions on some products such
} as asbestos powder and asbestos cloth. Unfortunately the legislators
} won't review thier laws once they are passed and the western world
} is spending needless billions of dollars in asbestos abatement in
} every old building that is renovated.
}
} Gordon
} Gordon Couger gcc-at-couger.com
}
} I collect links on information related to light microscopes.
} http://www.couger.com/microscope/links/gclinks.html
} Please forward any links or information you think might be useful to
} others.
} Microscope Manual at www.science-info.org
}
} } From: "Chiphead" {chiphead-at-sbcglobal.net}
}
} :
} : Warning: Non scientific answer:
} :
} : I remember an edition of "This Old House" or one of those types of
} shows
} : where they took some shingle samples to a lab for testing.
} :
} : The lab used polarized microscopy to do at least an "initial"
} screening
} : and determine that the shingles "most likely" contained asbestos.
} :
} : The "would do some additional testing" to verify (it was, and the
} house
} : needed to have a tent built around it, etc., etc.). I assume the
} : additional testing was something along the lines of TEM, etc.
} :
} : The conclusion I drew was that if you had materials that you were
} : expecting asbestos, this (polarized examination) was a good
} screening
} : tool. If the fibers didn't change from one specific color to
} another as
} : the analyzer was rotated, you didn't have asbestos. If they did,
} you
} : probably have asbestos, but there are other fibers that may do the
} same
} : thing.
} :
} : These were also fibers, not dust, and I don't know if that would
} affect
} : the results or not.
} :
} : Sorry if this wasn't useful.
} :
} : John R.
} :
} : -----Original Message-----
} : } From: by way of MicroscopyListserver [mailto:antuni-at-aol.com]
} : Sent: Saturday, May 22, 2004 9:23 AM
} : To: microscopy-at-ns.microscopy.com
} : Subject: [Microscopy] viaWWW: Identifying Asbestos Using ptical
} : Microscopy
} :
} :
} :
} : ------------------------------------------------------------------
} ------
} : ------
} : The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} : To Subscribe/Unsubscribe --
} : http://www.msa.microscopy.com/MicroscopyListserver
} : On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} : ------------------------------------------------------------------
} ------
} : -------
} :
} : Below is the result of your feedback form (NJZFM-ultra-55). It
} was
} : submitted by (antuni-at-aol.com) from
} : http://microscopy.com/MicroscopyListserver/MLFormMail.html on
} Friday,
} : May 21, 2004 at 16:19:26
} : ------------------------------------------------------------------
} ------
} : ---
} :
} : Email: antuni-at-aol.com
} : Name: Anthony Ribaudo
} :
} : Organization: Consultant
} :
} : Title-Subject: [Microscopy] [Filtered] Identifying Asbestos
} Using ptical
} : Microscopy
} :
} : Question: Can one identify asbestos only utilizing optical
} microscopy
} : without the aid of electron diffraction/TEM ?
} :
} :
} : Anthony Ribaudo
} :
} :
} :
} : ------------------------------------------------------------------
} ------
} : ---
} :
} :
} :
} :
} :
}
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Mon May 24 09:50:34 2004



From: Philip Oshel :      peoshel-at-wisc.edu
Date: Sun, 23 May 2004 16:17:00 -0500
Subject: [Microscopy] Gatan cryoholder -- forgot the 'scope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Listers,

Sorry, I forgot to state that the cryoholder we're looking for would
be for a Hitachi S-900 in-lens FESEM, which uses a Hitachi TEM holder.

Phil
--
Philip Oshel
Supervisor, AMFSC and BBPIC microscopy facilities
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706 - 1284
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)


From MicroscopyL-request-at-ns.microscopy.com Mon May 24 11:14:18 2004



From: Bard, Timothy J. :      Timothy.Bard-at-SYLVANIA.com
Date: Mon, 24 May 2004 12:26:01 -0400
Subject: [Microscopy] Re: RE: viaWWW: Identifying Asbestos Using ptical

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Gordon,

It is hard to believe other forms of asbestos: amosite asbestos(grunerite), actinolite-tremolite asbestos, anthophyllite (asbestos)and crocidolite asbestos(riebeckite) are not harmful. Have any studies been done on this? Chrysotile asbestos makes up to +90% of all asbestos in North America. (please correct if somebody has published figures on this).

Asbestos has been linked to mesothelioma and, also, asbestos is the cause of asbestosis(also harmful to humans). Isn't it?

I agree. If asbestos studies had been fully reviewed we would still using asbestos it has unique chemical and physical properties, especially for a mineral. The politicians used asbestos as a political football first scaring the public then passing legislation covering asbestos in schools, AHERA. Yes many problems have occurred as a result of this. The regulation should be reviewed and changed(if if hasn't already been).

Yes, asbestos can be identified by polarized light microscopy, not OLM or PCM. Using dispersion staining techniques to determine refractive indices, a properly prepared asbestos fiber can be identified, by PLM. Other measured properties: pleochroism, sign of elongation, and angle of extinction plus physical characteristic further aid the PLM analyst in the proper identification. Preparation of the fiber and proper training of the analyst are the critical factors here. Binder/matrix effects may hinder the analysis, for example chrysotile in vinyl floor tile.

Where are all the McCrone(ies) on this? Maybe this subject has been hashed out in the past. If so pardon me for the repeated redundancy.

Regards,

Timothy J. Bard, Scientist
Scanning Electron Microscopy
OSRAM SYLVANIA
Towanda, PA

U




-----Original Message-----
} From: Gordon Couger [mailto:gcouger-at-provalue.net]
Sent: Saturday, May 22, 2004 5:09 PM
To: Chiphead; 'by way of MicroscopyListserver';
microscopy-at-ns.microscopy.com

Only one of the forms of asbestos is harmful to humans. Chrysotile
and then only if it is inhaled and the particle size is such it
lodges in the lung so it can form a site for cancer to get a start.
Chrysotile is only a small percentage of asbestos and the particle
size must fall in the 5 to 15 micron range to stay in the lungs.

These conditions are almost only found in asbestos worker and miners
and not in the general public. In the case of high exposure to
asbestos cigarette smoking is also a very large contributor to the
cancer caused by asbestos.

Had the asbestos studies been allowed to be fully peer reviewed
before legislation was passed we would probably still be using
asbestos in many applications today with more protection for
asbestos workers and minors and restrictions on some products such
as asbestos powder and asbestos cloth. Unfortunately the legislators
won't review their laws once they are passed and the western world
is spending needless billions of dollars in asbestos abatement in
every old building that is renovated.

Gordon
Gordon Couger gcc-at-couger.com

I collect links on information related to light microscopes.
http://www.couger.com/microscope/links/gclinks.html
Please forward any links or information you think might be useful to
others.
Microscope Manual at www.science-info.org

} From: "Chiphead" {chiphead-at-sbcglobal.net}

:
: Warning: Non scientific answer:
:
: I remember an edition of "This Old House" or one of those types of
shows
: where they took some shingle samples to a lab for testing.
:
: The lab used polarized microscopy to do at least an "initial"
screening
: and determine that the shingles "most likely" contained asbestos.
:
: The "would do some additional testing" to verify (it was, and the
house
: needed to have a tent built around it, etc., etc.). I assume the
: additional testing was something along the lines of TEM, etc.
:
: The conclusion I drew was that if you had materials that you were
: expecting asbestos, this (polarized examination) was a good
screening
: tool. If the fibers didn't change from one specific color to
another as
: the analyzer was rotated, you didn't have asbestos. If they did,
you
: probably have asbestos, but there are other fibers that may do the
same
: thing.
:
: These were also fibers, not dust, and I don't know if that would
affect
: the results or not.
:
: Sorry if this wasn't useful.
:
: John R.
:
: -----Original Message-----
: } From: by way of MicroscopyListserver [mailto:antuni-at-aol.com]
: Sent: Saturday, May 22, 2004 9:23 AM
: To: microscopy-at-ns.microscopy.com
: Subject: [Microscopy] viaWWW: Identifying Asbestos Using ptical
: Microscopy
:
:
:
: ------------------------------------------------------------------
------
: ------
: The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
: To Subscribe/Unsubscribe --
: http://www.msa.microscopy.com/MicroscopyListserver
: On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
: ------------------------------------------------------------------
------
: -------
:
: Below is the result of your feedback form (NJZFM-ultra-55). It
was
: submitted by (antuni-at-aol.com) from
: http://microscopy.com/MicroscopyListserver/MLFormMail.html on
Friday,
: May 21, 2004 at 16:19:26
: ------------------------------------------------------------------
------
: ---
:
: Email: antuni-at-aol.com
: Name: Anthony Ribaudo
:
: Organization: Consultant
:
: Title-Subject: [Microscopy] [Filtered] Identifying Asbestos
Using ptical
: Microscopy
:
: Question: Can one identify asbestos only utilizing optical
microscopy
: without the aid of electron diffraction/TEM ?
:
:
: Anthony Ribaudo
:
:
:
: ------------------------------------------------------------------
------
: ---
:
:
:
:
:






From MicroscopyL-request-at-ns.microscopy.com Mon May 24 12:48:22 2004



From: Stacey Andringa :      Stacey.Andringa-at-uc.edu
Date: Mon, 24 May 2004 14:00:47 -0400
Subject: [Microscopy] immunohistology and fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have a student in my lab who wants to do immunohist on mouse skin. He
is looking at RPA. His current protocol for RPA is on cells. He was
able to get skin from mice treated with BPDE, he fixed the small strips
of skin in 4% paraformaldehyde in Dulbecco's PBS. He fixed it for 20-24
hours, then put it in 70% ethanol followed by routine paraffin
embedding.
Was the fixation time too long? I always fixed for 4-6 hours tops.
Will this affect his immunohist?
Thanks for any help or advice on this.

Stacey Andringa



From MicroscopyL-request-at-ns.microscopy.com Mon May 24 12:52:09 2004



From: Stacey Andringa :      Stacey.Andringa-at-uc.edu
Date: Mon, 24 May 2004 14:04:36 -0400
Subject: [Microscopy] cicadas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I work in Cincinnati and we are experiencing our 17 year cicada
invasion. I have never worked with insects and I am interested in
fixing some for light microscopy. What fixative would work on them and
how do I fix them - remove wings, pierce abdomen, etc.? This is for fun
as much as education.
Thanks for your help.

Stacey Andringa



From MicroscopyL-request-at-ns.microscopy.com Mon May 24 12:53:49 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 24 May 2004 11:11:24 -0700
Subject: [Microscopy] Re: HRTEM: how resolutions defined in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 22, 2004, at 3:07 AM, Juha Dapper wrote:

} For a FEG-HRTEM, what is the "line resolution" as
} opposed to the "information resolution"? What is the
} difference between each other?
}
} How could I achieve the "line resolution" and the
} "information resolution" in TEM experiments? And, how
} can I resolve (identify, or "see") the "line
} resolution" and the "information resolution"
} respectively from a state-of-the-art HREM image?
}
Dear Juha,
The line resolution is the closest distance between two edges that can
be resolved, and it can be pretty confusing. A better measure is point
resolution, which can be determined from the first zero of the contrast
transfer function at Shertzer focus. The information limit is the
farthest extent of the Thon rings that can be achieved. To determine
the point resolution, put a thin amorphous carbon specimen in the
scope, adjust the height to eucentric height, make sure the scope is
aligned as well as possible, go to a sufficiently high magnification so
that the spatial frequency of the expected point resolution is larger
than two--or better three--pixels (on a CCD, the actual elements, on
film, the grain separation or resolution of a scanner), then take an
image. Determine the power spectrum by taking the FFT of the
image--easy on a CCD; on film, scan with a high-resolution scanner,
then take the FFT. The spatial frequency of the first zero is the
point resolution of the image. The better the scope is adjusted and
the better the area of the specimen, the closer you will come to the
true resolution of the scope. The information limit is best seen by
shifting the image during the exposure. This produces Young's fringes
in the power spectrum, which makes the extent of the Thon rings easier
to see. The service engineers who set up our scopes used a switch
attached to a shift coil to do this test. In order to achieve the best
resolution and information transfer possible, make sure your scope is
as well aligned as possible. If your specimen is too thick, or
otherwise unsuitable, you may not achieve the specified performance, so
don't blame the equipment in this case. Good luck.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon May 24 13:59:59 2004



From: Frank.Karl-at-degussa.com
Date: Mon, 24 May 2004 15:11:08 -0400
Subject: [Microscopy] Re: RE: viaWWW: Identifying Asbestos Using ptical

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


OK, you sucked me in to asbestos and so here goes...........

Almost all of the minerals in existence were identified and
characterized by PLM at one time. Not all of these mineral were done as
well as the rest but still PLM is a powerful tool in the hands of an
experienced and trained microscopist. All of the legally defined asbestos
can be identified by PLM. Dispersion staining is a very useful tool for
this purpose. Vinyl tile, mortar, ceiling tiles, fluffy insulation, wall
plaster, and black or gray or green mastic - sample prep remains the
kingpin of all microscopical analysis just as it is in all analytical
testing.

My understanding is the current alert limits were based on measurements
made in asbestos mines and mills and extended to "normal" areas. This
level may be too low and in my opinion is a result of political games. It
seems some people have extreme sensitivity to asbestos and develop terminal
conditions at very low exposure levels. It is also my understanding these
studies were originally based on adults who were exposed as young adults
and older. I have not seen studies bases on exposure of children to
asbestos. So were do we draw the line? Yes, smoking makes it 400X more
likely to develop health complications (at least that was the number quote
when I did my asbestos certification training in Ohio. Not everyone who
was exposed to high levels developed health complications. I had a boss
who crawled through asbestos rewiring naval ships as summer work. He was a
long term chain smoker, and he lived into his 80's.

Government regulations created a industry of abatement and identification.
I have had the good fortune to participate at the identification and
quantification level for a number of years. The thing to remember is it
doesn't matter your personal beliefs or current scientific data when you're
compiling with a law, it is what the law says it is.







"Bard, Timothy
J." To: "Gordon Couger" {gcouger-at-provalue.net} ,
{Timothy.Bard-at-SYL "Microscopy (E-mail)" {microscopy-at-ns.microscopy.com}
VANIA.com} cc:
Subject: [Microscopy] RE: viaWWW: Identifying
05/24/2004 12:26 Asbestos Using optical Microscopy
PM








------------------------------------------------------------------------------

The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Gordon,

It is hard to believe other forms of asbestos: amosite asbestos(grunerite),
actinolite-tremolite asbestos, anthophyllite (asbestos)and crocidolite
asbestos(riebeckite) are not harmful. Have any studies been done on this?
Chrysotile asbestos makes up to +90% of all asbestos in North America.
(please correct if somebody has published figures on this).

Asbestos has been linked to mesothelioma and, also, asbestos is the cause
of asbestosis(also harmful to humans). Isn't it?

I agree. If asbestos studies had been fully reviewed we would still using
asbestos it has unique chemical and physical properties, especially for a
mineral. The politicians used asbestos as a political football first
scaring the public then passing legislation covering asbestos in schools,
AHERA. Yes many problems have occurred as a result of this. The
regulation should be reviewed and changed(if if hasn't already been).

Yes, asbestos can be identified by polarized light microscopy, not OLM or
PCM. Using dispersion staining techniques to determine refractive indices,
a properly prepared asbestos fiber can be identified, by PLM. Other
measured properties: pleochroism, sign of elongation, and angle of
extinction plus physical characteristic further aid the PLM analyst in the
proper identification. Preparation of the fiber and proper training of the
analyst are the critical factors here. Binder/matrix effects may hinder
the analysis, for example chrysotile in vinyl floor tile.

Where are all the McCrone(ies) on this? Maybe this subject has been hashed
out in the past. If so pardon me for the repeated redundancy.

Regards,

Timothy J. Bard, Scientist
Scanning Electron Microscopy
OSRAM SYLVANIA
Towanda, PA

U




-----Original Message-----
} From: Gordon Couger [mailto:gcouger-at-provalue.net]
Sent: Saturday, May 22, 2004 5:09 PM
To: Chiphead; 'by way of MicroscopyListserver';
microscopy-at-ns.microscopy.com


Only one of the forms of asbestos is harmful to humans. Chrysotile
and then only if it is inhaled and the particle size is such it
lodges in the lung so it can form a site for cancer to get a start.
Chrysotile is only a small percentage of asbestos and the particle
size must fall in the 5 to 15 micron range to stay in the lungs.

These conditions are almost only found in asbestos worker and miners
and not in the general public. In the case of high exposure to
asbestos cigarette smoking is also a very large contributor to the
cancer caused by asbestos.

Had the asbestos studies been allowed to be fully peer reviewed
before legislation was passed we would probably still be using
asbestos in many applications today with more protection for
asbestos workers and minors and restrictions on some products such
as asbestos powder and asbestos cloth. Unfortunately the legislators
won't review their laws once they are passed and the western world
is spending needless billions of dollars in asbestos abatement in
every old building that is renovated.

Gordon
Gordon Couger gcc-at-couger.com

I collect links on information related to light microscopes.
http://www.couger.com/microscope/links/gclinks.html
Please forward any links or information you think might be useful to
others.
Microscope Manual at www.science-info.org

} From: "Chiphead" {chiphead-at-sbcglobal.net}

:
: Warning: Non scientific answer:
:
: I remember an edition of "This Old House" or one of those types of
shows
: where they took some shingle samples to a lab for testing.
:
: The lab used polarized microscopy to do at least an "initial"
screening
: and determine that the shingles "most likely" contained asbestos.
:
: The "would do some additional testing" to verify (it was, and the
house
: needed to have a tent built around it, etc., etc.). I assume the
: additional testing was something along the lines of TEM, etc.
:
: The conclusion I drew was that if you had materials that you were
: expecting asbestos, this (polarized examination) was a good
screening
: tool. If the fibers didn't change from one specific color to
another as
: the analyzer was rotated, you didn't have asbestos. If they did,
you
: probably have asbestos, but there are other fibers that may do the
same
: thing.
:
: These were also fibers, not dust, and I don't know if that would
affect
: the results or not.
:
: Sorry if this wasn't useful.
:
: John R.
:
: -----Original Message-----
: } From: by way of MicroscopyListserver [mailto:antuni-at-aol.com]
: Sent: Saturday, May 22, 2004 9:23 AM
: To: microscopy-at-ns.microscopy.com
: Subject: [Microscopy] viaWWW: Identifying Asbestos Using ptical
: Microscopy
:
:
:
: ------------------------------------------------------------------
------
: ------
: The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
: To Subscribe/Unsubscribe --
: http://www.msa.microscopy.com/MicroscopyListserver
: On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
: ------------------------------------------------------------------
------
: -------
:
: Below is the result of your feedback form (NJZFM-ultra-55). It
was
: submitted by (antuni-at-aol.com) from
: http://microscopy.com/MicroscopyListserver/MLFormMail.html on
Friday,
: May 21, 2004 at 16:19:26
: ------------------------------------------------------------------
------
: ---
:
: Email: antuni-at-aol.com
: Name: Anthony Ribaudo
:
: Organization: Consultant
:
: Title-Subject: [Microscopy] [Filtered] Identifying Asbestos
Using ptical
: Microscopy
:
: Question: Can one identify asbestos only utilizing optical
microscopy
: without the aid of electron diffraction/TEM ?
:
:
: Anthony Ribaudo
:
:
:
: ------------------------------------------------------------------
------
: ---
:
:
:
:
:











From MicroscopyL-request-at-ns.microscopy.com Mon May 24 14:15:01 2004



From: Michael O'Keefe :      MAOKeefe-at-lbl.gov
Date: Mon, 24 May 2004 12:27:19 -0700
Subject: [Microscopy] Re: HRTEM: how resolutions defined in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Juha,

There are three commonly employed definitions of "HRTEM resolution". All depend on
how information is transferred by the objective lens from the specimen
"exit-surface wave" to the image intensity spectrum (the Fourier transform of the
image intensity).

(1) "Fringe" resolution, or "line" resolution (sometimes called "lattice-plane"
resolution), is measured from the highest spatial frequency that is present in the
image intensity spectrum (and is thus detectable in the optical diffractogram). It
may include non-physical detail, since it may include components from a
"second-order" or "non-linear" interference generating a half-spacing term with
spatial frequency up to twice that of the highest-frequency diffracted beam passed
by the objective aperture and the convergence cross-coefficient envelope
["Resolution-damping functions in non-linear images", M.A. O'Keefe, in 37th Ann.
Proc. EMSA, San Antonio, Texas (1979) 556-557]. For a properly aligned microscope,
it is not affected by microscope spread-of-focus and depends only on factors such
as vibration or detector MTF.

(2) "Linear-image" resolution, or "information-limit" resolution, is measured by
the highest spatial frequency transferred linearly from the amplitude spectrum (the
specimen "exit-surface wave") to the image intensity spectrum with any old phase.
Transferred frequencies fall within one or more passbands, but other (lower)
frequencies may be blocked. Increased underfocus will push the convergence-limited
damping function to higher frequencies, thus this limit (often called just the
"information limit") to point resolution is a function of microscope
spread-of-focus. Spread of focus is a function of objective lens chromatic
aberration and variations in lens current and in electron beam energy (and of
vertical vibration of the specimen within the lens).

(3) "Scherzer" resolution, or "structure-image" resolution (sometimes called
"point" resolution or "point-to-point" resolution) is measured by the highest
linearly-transferred spatial
frequency that can be passed when no lower frequencies are blocked or passed with
opposite phase. The Scherzer image is important because it is (an approximation
to) a projection of the specimen structure (to a limited resolution) in the
direction of the incident beam. For images obtained at the Scherzer "optimum"
defocus, the Scherzer resolution is generally defined by the upper limit of the
low-frequency same-phase passband. Scherzer resolution is a function of objective
lens spherical aberration and electron wavelength.

FEG-HRTEMs can transfer spatial frequencies out to quite high values (their
information limits are much better than their Scherzer resolutions). Since we
know how linear transfer varies with objective lens defocus, we can use a series of
images to produce a single reconstructed image containing correctly-phased
components all the way out to the microscope information limit [see work by Van
Dyck, Coene and Thust]. The latest issue of Microscopy Today shows an example of a
reconstructed image with 0.78 Angstrom resolution, although the microscope used has
a Scherzer resolution of only 1.7 Angstrom.

While the phases of linearly-transferred image components can be corrected so they
can be used to extend resolution out to the microscope information limit, the
non-lnear components that create the "fringe resolution" usually transfer no new
information. These "cross-aperture" components can be frequency-doubled -- for
example, a 200 beam from the specimen may interfere with a -200 beam to create a
400 image component. Then we will see spacings of a/4 in the image even though no
information about the a/4 spacing was transferred from the specimen. If we managed
to "dis-entangle" the image's non-linear 400 component with its a/4 spacing, it
would still only provide us with information on a/2 spacings in the specimen.

So the short answer to your question is -- the difference between "line resolution"
as opposed to the "information resolution" is that an image can contain specimen
information out to the "information resolution", but any finer image spacings --
out to the "line resolution" -- are not new information about the specimen, but
were generated by non-linear interferences in the objective lens.

Mike O'Keefe


Juha Dapper wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Dear microscopist experts,
}
} When I read literatures, I have been all the time
} confused by the resolution definition in transmission
} electron microscopy. Could you please kindly help?
}
} For a FEG-HRTEM, what is the "line resolution" as
} opposed to the "information resolution"? What is the
} difference between each other?
}
} How could I achieve the "line resolution" and the
} "information resolution" in TEM experiments? And, how
} can I resolve (identify, or "see") the "line
} resolution" and the "information resolution"
} respectively from a state-of-the-art HREM image?
}
} Thank you for your time. Wish you a nice weekend,
}
} -Juha
}
}
}
}
} __________________________________
} Do you Yahoo!?
} Yahoo! Domains – Claim yours for only $14.70/year
} http://smallbusiness.promotions.yahoo.com/offer



From MicroscopyL-request-at-ns.microscopy.com Mon May 24 15:36:21 2004



From: Pat Connelly :      psconnel-at-sas.upenn.edu
Date: Mon, 24 May 2004 16:48:12 -0400
Subject: [Microscopy] Haskris water chiller/recycler

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have a water chiller from 1996 (Haskris RO75 with B H M + T options).

On checking the water level today, I noticed that the water was a bit green and
hence needed to be changed.
I drained the water storage tank, removed the nylon strainer,
replaced it with a clean spare
and as luck would have it I also needed to change the drain hose.

From experience I knew that when I first change the water, there is
normally a rinsing out
of the system and a noticeable amount of algae (? greenish gunk)
comes out of the discharge
pipe and I change the water another time before adding the
anti-algecid and closing up the system.

Today I got some air into the system and when I turned it back on the
water turned so turbid
that I could not see the bottom coils in the tank. I repeated the
draining/filling cycle at least
8 times until there was only a slight greenish cast to the water and
I gave up until tomorrow.

Does anyone have any idea why this would happen now and not before?
Is there any way that I can be assured that I have removed as much of
the algae from the
system as I can, other than to keep changing the water as long as I
can see small clumps
coming out into the water tank?

Thanks in advance,
Pat Connelly
Dept. of Biology
Univ. of PA
Philadelphia, PA 19104


From MicroscopyL-request-at-ns.microscopy.com Mon May 24 15:48:37 2004



From: Libby Shaw :      elshaw-at-MIT.EDU
Date: Mon, 24 May 2004 17:00:32 -0400
Subject: [Microscopy] RE: viaWWW: Identifying Asbestos Using optical Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi, Timothy:

I seem to recollect that Neal Rowlands, who was Shared Experimental
Facilities Manager in our Center some years ago, had done some asbestos
research at McGill U. along the lines of your question. I recall him
saying that the difference in lung damage caused by different forms of
asbestos had to do with the difference in crystal shape. The tiny-needle
shape was the nastiest.

Best regards,

Libby Shaw


} Subject: [Microscopy] RE: viaWWW: Identifying Asbestos Using optical
} Microscopy
} Date: Mon, 24 May 2004 12:26:01 -0400
} From: "Bard, Timothy J." {Timothy.Bard-at-SYLVANIA.com}
} To: "Gordon Couger" {gcouger-at-provalue.net} ,
} "Microscopy (E-mail)" {microscopy-at-ns.microscopy.com}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Address: MIT Room 13-4149 Tel: 617-253-5045
77 Massachusetts Avenue Email: elshaw-at-mit.edu
Cambridge, MA 02139 Fax: 617-258-6478
http://web.mit.edu/cmse/www/
****************************************************************


From MicroscopyL-request-at-ns.microscopy.com Tue May 25 01:05:16 2004



From: John Shields :      jpshield-at-uga.edu
Date: Mon, 24 May 2004 17:46:40 -0400
Subject: [Microscopy] M&M 2004 student volunteers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

To Students attending M&M2004 in Savannah,
If you are a student that will be attending (or would like
to the attend) the M&M meeting in Savannah this summer, we
would appreciate your help by assisting at the sessions.
You would be volunteering as a "projectionist" for various
sessions, most likely ones you would attend anyway. There
is a training session Sunday afternoon (details when you
volunteer).

What are the benefits of volunteering? Volunteers will be
given free registration--a badge, opening night reception
ticket and a CD of the abstracts.

Who to contact:
John Shields (see address below) at jshields-at-cb.uga.edu


--
John Shields
E.M. Lab, 151 Barrow Hall
University of Georgia
Athens, GA 30602
706-542-4080



From MicroscopyL-request-at-ns.microscopy.com Tue May 25 03:13:08 2004



From: Philip Koeck :      Philip.Koeck-at-biosci.ki.se
Date: Tue, 25 May 2004 10:24:45 +0200
Subject: [Microscopy] RE: Re: HRTEM: how resolutions defined in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,

I just want to get this clear and look at the consequences of what you
write.

If a nonlinear component of frequency 2f is present in an image, will it
have been damped by the ctf-envelope just like the corresponding linear
component of frequency f? In that case the presence of a high-resolution
component doesn't say much about the quality of a microscope or a
micrograph as long as it is unclear whether the component is linear or
frequency doubled.
Now for the practical consequences:
With what sort of specimens would you expect frequency doubling? Could
it happen with protein crystals (10-20 nm thick) using a 200 or 300 kV
microscope?

Philip

-----Original Message-----
} From: Michael O'Keefe [mailto:MAOKeefe-at-lbl.gov]

(1) "Fringe" resolution, or "line" resolution (sometimes called
"lattice-plane" resolution), is measured from the highest spatial
frequency that is present in the image intensity spectrum (and is thus
detectable in the optical diffractogram). It may include non-physical
detail, since it may include components from a "second-order" or
"non-linear" interference generating a half-spacing term with spatial
frequency up to twice that of the highest-frequency diffracted beam
passed by the objective aperture and the convergence cross-coefficient
envelope

While the phases of linearly-transferred image components can be
corrected so they can be used to extend resolution out to the microscope
information limit, the non-linear components that create the "fringe
resolution" usually transfer no new information. These "cross-aperture"
components can be frequency-doubled -- for example, a 200 beam from the
specimen may interfere with a -200 beam to create a 400 image component.
Then we will see spacings of a/4 in the image even though no information
about the a/4 spacing was transferred from the specimen. If we managed
to "dis-entangle" the image's non-linear 400 component with its a/4
spacing, it would still only provide us with information on a/2 spacings
in the specimen.

So the short answer to your question is -- the difference between "line
resolution" as opposed to the "information resolution" is that an image
can contain specimen information out to the "information resolution",
but any finer image spacings -- out to the "line resolution" -- are not
new information about the specimen, but were generated by non-linear
interferences in the objective lens.






From MicroscopyL-request-at-ns.microscopy.com Tue May 25 05:02:13 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 24 May 2004 15:24:38 -0700
Subject: [Microscopy] Re: Haskris water chiller/recycler

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 24, 2004, at 1:48 PM, Pat Connelly wrote:

} From experience I knew that when I first change the water, there is
} normally a rinsing out
} of the system and a noticeable amount of algae (? greenish gunk) comes
} out of the discharge
} pipe and I change the water another time before adding the
} anti-algecid and closing up the system.
}
Dear Pat,
Are you sure that the gunk is algae and not Cu oxide? If the latter,
you should consider adding a corrosion inhibitor to the water.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Tue May 25 06:08:07 2004



From: Philip Koeck :      Philip.Koeck-at-biosci.ki.se
Date: Tue, 25 May 2004 13:11:29 +0200
Subject: [Microscopy] RE: Microscopy: Why FEG?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I thought some people might be interested in a follow-up of this
discussion.
(Don't tell me if your not, just ignore me.)

I've taken some images of carbon film on our CM120 with a brand new
LaB6-filament just to get an idea of the limitations of this microscope.

I get information limits between 15 and 5 Angstrom depending on
magnification and defocus (only "biological" defocus values) and the
information limit does not improve with decreasing magnification as
suggested by CTF-explorer. In fact the opposite is true (at least for
constant exposure levels). Some similar data for a FEG would be
interesting.

One thing is clear: Reaching 10 Angstrom and slightly better should not
be a problem with a well maintained LaB6 microscope.

For the details see the link "reference data for the CM120" at
http://www.csb.ki.se/users/philip/philown.html. This is a big word file
so you have to be patient when you download it.

Philip




From MicroscopyL-request-at-ns.microscopy.com Tue May 25 06:31:39 2004



From: David_Bell-at-millipore.com
Date: Tue, 25 May 2004 07:43:16 -0400
Subject: [Microscopy] Re: [Detected by Millipore as possible Spam] Re: Haskris

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Bill,

That's exactly what I was thinking, as well. Do you have any good
recommendations for use as a corrosion inhibitor?

Thanks,

David



David Bell
Scientist
Electron Microscopy Lab
Millipore Corporation
80 Ashby Road
Bedford, MA 01730
(781) 533-2108






Bill Tivol {tivol-at-caltech.edu}
05/24/2004 06:24 PM


To: microscopy-at-msa.microscopy.com
cc:
Subject: [Microscopy] [Detected by Millipore as possible Spam] Re: Haskris water
chiller/recycler




------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


On May 24, 2004, at 1:48 PM, Pat Connelly wrote:

} From experience I knew that when I first change the water, there is
} normally a rinsing out
} of the system and a noticeable amount of algae (? greenish gunk) comes
} out of the discharge
} pipe and I change the water another time before adding the
} anti-algecid and closing up the system.
}
Dear Pat,
Are you sure that the gunk is algae and not Cu oxide? If
the latter,
you should consider adding a corrosion inhibitor to the water.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu








From MicroscopyL-request-at-ns.microscopy.com Tue May 25 07:53:51 2004



From: Tony Greco :      tgreco-at-seas.marine.usf.edu
Date: Tue, 25 May 2004 08:57:38 -0400
Subject: [Microscopy] Haskris chiller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Bill
I'm pretty sure I have Cu oxide along the walls of my Haskris
chiller as it won't come off. What corrosion inhibitor do you suggest
and how do I remove the Cu oxide already present?

Tony Greco

--
Anthony M. Greco
Electron Microscope Manager
College of Marine Science
University of South Florida
140 7th Avenue S.
St. Petersburg, Fl 33701

voice: (727) 553-1595
email: tgreco-at-marine.usf.edu




From MicroscopyL-request-at-ns.microscopy.com Tue May 25 08:53:39 2004



From: Sherwood, Margaret :      MSHERWOOD-at-PARTNERS.ORG
Date: Tue, 25 May 2004 10:00:51 -0400
Subject: [Microscopy] Re: [Detected by Millipore as possible Spam] Re:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I'm glad that this issue is still being addressed. I have a Haskris chiller
(since 1988) which services our Philips CM10 TEM. We continually have this
"green algae" in the water and which seems to collect around the filter unit
in the system. I have been concerned about this in the past, and asked our
service engineer, but it doesn't appear to impede the flow of water or the
operation of the chiller. (We have had other minor problems with the
chiller that B & G has fixed, but nothing connected to the water). I was
interested to hear about the possibility of it due to Cu oxide. I would
also like to know what corrosion inhibitor can be added to the water.

Peggy

Peggy Sherwood
Lab Associate, Photopathology
Wellman Laboratories of Photomedicine (W224)
Massachusetts General Hospital
55 Fruit Street
Boston, MA 02114
617-724-4839 (voice mail)
617-726-6983 (lab)
617-726-3192 (fax)
msherwood-at-partners.org


-----Original Message-----
} From: David_Bell-at-millipore.com [mailto:David_Bell-at-millipore.com]
Sent: Tuesday, May 25, 2004 7:43 AM
To: Bill Tivol
Cc: microscopy-at-msa.microscopy.com

Hi Bill,

That's exactly what I was thinking, as well. Do you have any good
recommendations for use as a corrosion inhibitor?

Thanks,

David



David Bell
Scientist
Electron Microscopy Lab
Millipore Corporation
80 Ashby Road
Bedford, MA 01730
(781) 533-2108






Bill Tivol {tivol-at-caltech.edu}
05/24/2004 06:24 PM


To: microscopy-at-msa.microscopy.com
cc:
Subject: [Microscopy] [Detected by Millipore as possible
Spam] Re: Haskris water
chiller/recycler




----------------------------------------------------------------------------
--
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


On May 24, 2004, at 1:48 PM, Pat Connelly wrote:

} From experience I knew that when I first change the water, there is
} normally a rinsing out
} of the system and a noticeable amount of algae (? greenish gunk) comes
} out of the discharge
} pipe and I change the water another time before adding the
} anti-algecid and closing up the system.
}
Dear Pat,
Are you sure that the gunk is algae and not Cu oxide? If
the latter,
you should consider adding a corrosion inhibitor to the water.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu








From MicroscopyL-request-at-ns.microscopy.com Tue May 25 08:56:31 2004



From: Juha Dapper :      juha_dapper-at-yahoo.com
Date: Tue, 25 May 2004 07:07:53 -0700 (PDT)
Subject: [Microscopy] Re: HRTEM: how resolutions defined in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thank you, Michael and Bill, for your kind reply. I
really appreciate your remarkable answers to my
questions.

After carefully reading your answers, I've got more
questions.

1. If the "line resolution" is so confusing and
untrustworthy, why TEM manufactures used to apply it
to claim the quality of a microscope?

2. If an HREM iamge was taken without a certain
objective aperture, how can we tell if a
high-frequency spacing present in the image (or the
corresponding spot in its FFT) is due to linear
transfer or just by "cross-aperture" non-linear
interference?

3. About the "non-linear interferences". Do they
happen already during the process of electron-specimen
interaction, or after the back-focal plane of the
objective lens during the electron wave propagation?

4. In Michael's email, it is said "Since we know how
linear transfer varies with objective lens defocus, we
can use a series of images to produce a single
reconstructed image containing correctly-phased
components all the way out to the microscope
information limit [see work by Van Dyck, Coene and
Thust]."

Did you mean that the reconstruction method using a
focal-series can only apply to the linear transfer
case, i.e. the case of a weak phase object? If not,
how the non-linear interference effects are treated?

5. In Bill's email, you mentioned that "The
information limit is best seen by shifting the image
during the exposure. This produces Young's fringes in
the power spectrum,".

Could you please tell me in more details about how to
"shift the image to produce Young's fringes"? This
may be very practical and helpful to me.

Again thank you very much for your instructions.

-juha




__________________________________
Do you Yahoo!?
Friends. Fun. Try the all-new Yahoo! Messenger.
http://messenger.yahoo.com/


From MicroscopyL-request-at-ns.microscopy.com Tue May 25 10:42:49 2004



From: psconnel-at-sas.upenn.edu
Date: Tue, 25 May 2004 11:54:27 -0400
Subject: [Microscopy] Re: Haskris water chiller/recycler

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society
of America On May 24, 2004, at 1:48 PM,
Pat Connelly {psconnel-at-sas.upenn.edu} wrote:
} } From experience I knew that when I first change the water,
there is normally a rinsing out of the system and
a noticeable amount of algae (? greenish gunk) comes out
of the discharge pipe and I change the water another time
before adding the anti-algecid and closing up the system. etc.
} }
} Dear Pat,
} Are you sure that the gunk is algae and not Cu oxide?
} If the latter, you should consider adding a corrosion
} inhibitor to the water.
} Bill Tivol, PhD
} EM Scientist and Manager
} Cryo-Electron Microscopy Facility
} Broad Center, Mail Code 114-96
} California Institute of Technology
} Pasadena CA 91125
} (626) 395-8833 tivol-at-caltech.edu
=======
Bill,
I took a sample from the fourth change of water and examined it
under a light microscope. I saw mostly tiny cellular clumps and
cellular sheets with a few very small, redish, spike like, crystals
embedded into them and that is why I thought of algae.
I do agree that the green color most likely comes from the copper
because there is no light in the system for photosynthesis.
I have sprinkled dichlorophene onto the water each time that
I have changed it since the first fill-up. The label states that
it is both a fungicide and a bactericide.
As the water supply, I use reverse osmosis water and some
Philadelphia tap water so that there are some ions.
The ph of both sources is usually acid, running about pH 5.5.
I did not check the pH of the water that I drained yesterday
but after rinning overnight the water resevoir is now also at
pH 5.5.

I am interested in your response as to a corrosion inhibitor.
Pat Connelly


From MicroscopyL-request-at-ns.microscopy.com Tue May 25 12:24:55 2004



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Tue, 25 May 2004 10:36:17 -0700
Subject: [Microscopy] asbestos risks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Rosemary;
My recollection from the asbestos work I did at Lockheed was
that most of the asbestos fibers could produce mesothelioma, but that in
North America, chrysotile is the predominant type of asbestos. Sources
of exposure vary, but the connection to mesothelioma was first noticed
after WWII when a significant population that also smoked heavily was
involved in refitting steam ships that had large amounts of asbestos
insulation. The death of actor Steve McQueen highlighted the fact that
even something like dirt bike riding in areas naturally high in asbestos
could lead to a high risk exposure.

John Mardinly
Intel

-----Original Message-----
} From: Rosemary White [mailto:Rosemary.White-at-csiro.au]
Sent: Sunday, May 23, 2004 4:28 PM
To: microscopy-at-msa.microscopy.com

This is off the topic of identifying asbestos, but thought I should
comment
that just yesterday there was a program on the radio (streaming audio
available at http://www.abc.net.au/rn/talks/bbing/, transcript available
by
Thursday) about new cases of asbestosis and mesothelioma showing up.
The
young people (20s and 30s) now getting this lived in high asbestos areas
as
kids, were children of builders (and builders themselves) who demolished
or
renovated fibro houses, etc. Perhaps just an Australian problem....
Rosemary

} From: "Gordon Couger" {gcouger-at-provalue.net}
} Date: Sat, 22 May 2004 16:09:06 -0500
} To: "Chiphead" {chiphead-at-sbcglobal.net} , "'by way of
MicroscopyListserver'"
} {antuni-at-aol.com} , {microscopy-at-ns.microscopy.com}
} Subject: [Microscopy] Re: RE: viaWWW: Identifying Asbestos Using
ptical
} Microscopy
}
}
}
}
------------------------------------------------------------------------
------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe --
} http://www.msa.microscopy.com/MicroscopyListserver
} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
------}
-
}
} Only one of the forms of asbestos is harmful to humans. Chrysotile
} and then only if it is inhaled and the particle size is such it
} lodges in the lung so it can form a site for cancer to get a start.
} Chrysotile is only a small percentage of asbestos and the particle
} size must fall in the 5 to 15 micron range to stay in the lungs.
}
} These conditions are almost only found in asbestos worker and miners
} and not in the general public. In the case of high exposure to
} asbestos cigarette smoking is also a very large contributor to the
} cancer caused by asbestos.
}
} Had the asbestos studies been allowed to be fully peer reviewed
} before legislation was passed we would probably still be using
} asbestos in many applications today with more protection for
} asbestos workers and minors and restrictions on some products such
} as asbestos powder and asbestos cloth. Unfortunately the legislators
} won't review thier laws once they are passed and the western world
} is spending needless billions of dollars in asbestos abatement in
} every old building that is renovated.
}
} Gordon
} Gordon Couger gcc-at-couger.com
}
} I collect links on information related to light microscopes.
} http://www.couger.com/microscope/links/gclinks.html
} Please forward any links or information you think might be useful to
} others.
} Microscope Manual at www.science-info.org
}
} } From: "Chiphead" {chiphead-at-sbcglobal.net}
}
} :
} : Warning: Non scientific answer:
} :
} : I remember an edition of "This Old House" or one of those types of
} shows
} : where they took some shingle samples to a lab for testing.
} :
} : The lab used polarized microscopy to do at least an "initial"
} screening
} : and determine that the shingles "most likely" contained asbestos.
} :
} : The "would do some additional testing" to verify (it was, and the
} house
} : needed to have a tent built around it, etc., etc.). I assume the
} : additional testing was something along the lines of TEM, etc.
} :
} : The conclusion I drew was that if you had materials that you were
} : expecting asbestos, this (polarized examination) was a good
} screening
} : tool. If the fibers didn't change from one specific color to
} another as
} : the analyzer was rotated, you didn't have asbestos. If they did,
} you
} : probably have asbestos, but there are other fibers that may do the
} same
} : thing.
} :
} : These were also fibers, not dust, and I don't know if that would
} affect
} : the results or not.
} :
} : Sorry if this wasn't useful.
} :
} : John R.
} :
} : -----Original Message-----
} : } From: by way of MicroscopyListserver [mailto:antuni-at-aol.com]
} : Sent: Saturday, May 22, 2004 9:23 AM
} : To: microscopy-at-ns.microscopy.com
} : Subject: [Microscopy] viaWWW: Identifying Asbestos Using
ptical
} : Microscopy
} :
} :
} :
} : ------------------------------------------------------------------
} ------
} : ------
} : The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} : To Subscribe/Unsubscribe --
} : http://www.msa.microscopy.com/MicroscopyListserver
} : On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} : ------------------------------------------------------------------
} ------
} : -------
} :
} : Below is the result of your feedback form (NJZFM-ultra-55). It
} was
} : submitted by (antuni-at-aol.com) from
} : http://microscopy.com/MicroscopyListserver/MLFormMail.html on
} Friday,
} : May 21, 2004 at 16:19:26
} : ------------------------------------------------------------------
} ------
} : ---
} :
} : Email: antuni-at-aol.com
} : Name: Anthony Ribaudo
} :
} : Organization: Consultant
} :
} : Title-Subject: [Microscopy] [Filtered] Identifying Asbestos
} Using ptical
} : Microscopy
} :
} : Question: Can one identify asbestos only utilizing optical
} microscopy
} : without the aid of electron diffraction/TEM ?
} :
} :
} : Anthony Ribaudo
} :
} :
} :
} : ------------------------------------------------------------------
} ------
} : ---
} :
} :
} :
} :
} :
}
}
}
}







From MicroscopyL-request-at-ns.microscopy.com Tue May 25 12:56:12 2004



From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Tue, 25 May 2004 13:08:29 -0500
Subject: [Microscopy] Expiration standards for microscopy chemicals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Listers,

Does anyone else think it might be a useful exercise to try and
establish expiration dates for chemicals used in microscopy labs? I
understand that chemicals "expire" based upon what they are, what use
they are intended for, how they are used, what level of
accuracy/repeatability/etc. is expected from their use, how they are
stored, and many other factors. I have also read Rande Kline's
discussion of the problem in the July/August 2000 Microscopy Today on
the difficulties of establishing such dates (recommended).

The problem is that some work requires lab certification and strict
laboratory guidelines requiring that all chemicals be labeled with
expiration dates----whether or not those dates have any meaningful or
consistent basis in "reality". So far, I've been unable to find any
such set of standards and my strong impression is that labs set up their
own arbitrary chemical rotation regimen.

It might reasonably be suggested that MSA could be a clearinghouse for
establishing such standards for microscopy-related research. If enough
people think this is a useful project, I would be willing (please
somebody, stop me!!) to volunteer to put together a draft proposal. As
a starter, it would be useful to know how other labs handle this issue
and why.

Any thoughts?

Thanks,
Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility---We Do Small Well!
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.biotech.missouri.edu/emc/




From MicroscopyL-request-at-ns.microscopy.com Tue May 25 13:36:42 2004



From: Robert Simmons :      rsimmons-at-gsu.edu
Date: Tue, 25 May 2004 14:49:09 -0400
Subject: [Microscopy] Re: Re: Haskris water chiller/recycler

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I've had a similar problem with my chiller system, though mine got clogged
up pretty well before I figured out what was going on. The turbo pump in my
SEM was the culprit, I think. The water jacket on the turbo is aluminum and
there are lots of brass fittings in the Haskris and the heat sinks in the
SEM. I got a bunch of stuff that looked like green kitty litter out of the
system that proved to have high aluminum and copper levels in it. I think
there is some sort of an electrochemical gradient set up that causes some
sort of corrosion product to build up over time (I'm a biologist, not a
chemist). Small amounts of it look like green sand in the bottom of the
reservoir. I cleaned the bigger clogs out of the water jacket with
alternating dilute HCl and NaOH using a big syringe and Tygon tubing,
rinsing with DIW in between (Rube Goldberg lives on). It's been fine for a
few months now, but I keep an eye on the flow rates.

Robert Simmons

Dr Robert Simmons
Program Director
Biological Imaging Core Laboratory
Georgia State University
Atlanta, GA 30302-4010

404-651-3138
404-651-2509 FAX



From MicroscopyL-request-at-ns.microscopy.com Tue May 25 14:18:48 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Tue, 25 May 2004 12:36:24 -0700
Subject: [Microscopy] Re: [Detected by Millipore as possible Spam] Re: Haskris water chiller/recycler

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 25, 2004, at 4:43 AM, David_Bell-at-millipore.com wrote:

} That's exactly what I was thinking, as well. Do you have any good
} recommendations for use as a corrosion inhibitor?
}
Dear David,
When I was at Albany NY, we used a product called Aquatreet 42, a
Mo-based inhibitor, and, judging from measurements of the water flow
through the lenses each year, there was little corrosion build-up.
That product can be obtained in a minimum of 5 gal--enough to last a
lifetime--from Aqua Labs, Inc., P.O. Box 645, Amesbury MA 01913. I
just spoke recently with Tom Cass, (800) 343-0213, and they still sell
the product to people in your area. He also told me that Skasol makes
a comparable product for the West Coast people. David Marchman, (800)
839-1000, is the contact for Skasol, and Chris Killian, (310) 749-8807
is the technical person. I have no affiliation with Aqua except as a
satisfied customer, and none with Skasol except as a potentially
satisfied customer.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Tue May 25 14:25:29 2004



From: Steve Buckingham :      steve-at-kratos.com
Date: Tue, 25 May 2004 15:36:01 -0400
Subject: [Microscopy] Re: Haskris water chiller/recycler

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,
Try dissolving it in 10% HCl. If it disappears its a Cu compound.

Best Regards,

Steve

Steve Buckingham
Service Manager
Kratos Analytical Inc.,
100 Red Schoolhouse Road, Bldg A
Chestnut Ridge, NY 10977
ph (845) 426 6700 ext 202
fx (845) 818 {mailto:4095steve-at-kratos.com} 4095

steve-at-kratos.com




-----Original Message-----
} From: psconnel-at-sas.upenn.edu [mailto:psconnel-at-sas.upenn.edu]
Sent: Tuesday, May 25, 2004 11:54 AM
To: Bill Tivol
Cc: Microscopy-at-msa.microscopy.com

----------------------------------------------------------------------------
--
The Microscopy ListServer -- Sponsor: The Microscopy Society
of America On May 24, 2004, at 1:48 PM,
Pat Connelly {psconnel-at-sas.upenn.edu} wrote:
} } From experience I knew that when I first change the water,
there is normally a rinsing out of the system and
a noticeable amount of algae (? greenish gunk) comes out
of the discharge pipe and I change the water another time
before adding the anti-algecid and closing up the system. etc.
} }
} Dear Pat,
} Are you sure that the gunk is algae and not Cu oxide?
} If the latter, you should consider adding a corrosion
} inhibitor to the water.
} Bill Tivol, PhD
} EM Scientist and Manager
} Cryo-Electron Microscopy Facility
} Broad Center, Mail Code 114-96
} California Institute of Technology
} Pasadena CA 91125
} (626) 395-8833 tivol-at-caltech.edu
=======
Bill,
I took a sample from the fourth change of water and examined it
under a light microscope. I saw mostly tiny cellular clumps and
cellular sheets with a few very small, redish, spike like, crystals
embedded into them and that is why I thought of algae.
I do agree that the green color most likely comes from the copper
because there is no light in the system for photosynthesis.
I have sprinkled dichlorophene onto the water each time that
I have changed it since the first fill-up. The label states that
it is both a fungicide and a bactericide.
As the water supply, I use reverse osmosis water and some
Philadelphia tap water so that there are some ions.
The ph of both sources is usually acid, running about pH 5.5.
I did not check the pH of the water that I drained yesterday
but after rinning overnight the water resevoir is now also at
pH 5.5.

I am interested in your response as to a corrosion inhibitor.
Pat Connelly



From MicroscopyL-request-at-ns.microscopy.com Tue May 25 14:33:40 2004



From: Bill Tivol :      tivol-at-caltech.edu
Date: Tue, 25 May 2004 12:51:16 -0700
Subject: [Microscopy] Re: RE: Microscopy: Why FEG?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On May 25, 2004, at 4:11 AM, Philip Koeck wrote:

} I get information limits between 15 and 5 Angstrom depending on
} magnification and defocus (only "biological" defocus values) and the
} information limit does not improve with decreasing magnification as
} suggested by CTF-explorer. In fact the opposite is true (at least for
} constant exposure levels). Some similar data for a FEG would be
} interesting.
}
Dear Philip,
One of the acceptance tests for our Tecnai F30H was that we had to see
30 Thon rings at 2 um underfocus on a thin C specimen, which is
slightly beyond the 0.34 nm graphite spacing. We redo this test
periodically to be sure that the coherence of the beam is still up to
par. In fact, the service engineers were able to get many more Thon
rings than that. The spec on the info limit is 0.14 nm, but we can't
get Thon rings that far out at 2 um underfocus. To get the 30 Thon
rings, one has to find a good area of the C film, so we don't expect
that good a performance for biological specimens.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Tue May 25 15:42:30 2004



From: amcelwai-at-bucknell.edu (by way of MicroscopyListserver)
Date: Tue, 25 May 2004 15:55:09 -0500
Subject: [Microscopy] [Filtered] MicroscopyListserverviaWWW: Text/Manual on amphibian

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (amcelwai-at-bucknell.edu) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, May 25, 2004 at 14:55:05
---------------------------------------------------------------------------

Email: amcelwai-at-bucknell.edu
Name: Andrew

Organization: Bucknell University Biology Department

Title-Subject: [Microscopy] [Filtered] MListserver: New Text/Manual

Question: Hi all,

I am a graduate student at Bucknell University and I am currently studying amphibian physiology. My thesis project involves studying skeletal muscles and liver tissue using both transmission electron microscopy, and light microscopy. I am interested in learning about new techniques in microscopy, or histology and I was hoping that someone could recommend a book that covers curent methodologies and, or the best ways of doing histology/microscopy research. All of the literature I have is from the 70's and the 80's and some of my colleagues suggested to me that my sources of information are a tad bit outdated.

I would appreciate any feedback. Thanks.

-Andrew

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue May 25 17:22:31 2004



From: Therese D Brackenbury :      Therese.Brackenbury-at-anu.edu.au
Date: Wed, 26 May 2004 08:33:45 +1000
Subject: [Microscopy] nanoplast infiltration problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi everyone

I need some help.

I have used nanoplast resin for the first time following instructions
provided with the kit. When I came to examine the sections I found there
had been poor infiltration. I would welcome any suggestions from people
that have worked with the resin that could resolve this plus any other help
tips. Do I need to do a resin infiltration series similar to Araldite. I
did email Pelco (manufacturer of the resin) about this but so far not had a
response.

thanks Therese



From MicroscopyL-request-at-ns.microscopy.com Tue May 25 19:14:18 2004



From: Beauregard :      beaurega-at-westol.com
Date: Tue, 25 May 2004 20:25:00 -0400
Subject: [Microscopy] Re: Haskris chiller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Tony,

I had an Ice Wagon and a Nesslab chiller system. The green color you see
MIGHT BE algae. Everyone blames the green color on algae.

I know from chiller experience and being a chemist that copper pipes
corrode first to a black thin film of copper oxide material on the walls of
chiller copper pipes. Next a deposit of copper carbonate forms over the
initial copper oxide coating and on PE cartridge filters. The black color
only forms on copper surfaces.

Green deposits in chiller lines in my lab were always thicker AFTER going
through a diffusion pump. We had whole house filters before the DPs and
they collected a green scale. The lines between the filter and the DP
inlet were relatively clear of green deposits. Seems backwards, doesn't
it? FYI.

TESTING:
Remove the scale and look under a microscope as you add dilute HCl. If it
bubbles, it's most likely lime green copper carbonate, as I had. To find
out if it was copper and carboante, I would use the regular green water to
test for copper ions first. Take the green chiller solution and add HCl.
It will dissolve the turbidity and form a nice clear green solution. Add
enough ammonium hydroxide to make the solution fairly basic. If copper
ions are present, the solution will turn deep blue. Algae do not do this.
Next add more HCL and the solution will turn green again. If your solution
does this, it's copper carbonate. If you would like to reconfirm the blue
color again, you can repeat the NH4OH treatment on the same sample of
water. This reversible color change is not possible with algae. IMHO.

To test for carbonate, add some HCl to a new sample to just barely get a
clear solution. Adjust the pH to 7-9 with ammonium hydroxide. Don't use
NaOH as it has carbonate in it. Add some clear barium chloride solution to
this basic solution. Enough carbonate ions will remain dissolved to cause
a white precipitate of BaCO3 to form. That probably confirms the gas was
originally CO2. If you then add HCl, the white barium carbonate
precipitate disappears to form carbonic acid and barium chloride. This
chemical behavior confirms CO3 for sure.

If you find the problem is algae, then you can add a small amount of a
quaternary ammonium salt. This extremely long named chemical will end with
quaternary ammonium salt and can be bought at any swimming pool supply as
an algicide. Estimate the volume of your system and add the recommended
amount. As I recall, a quart treats about 5,000 to 10,000 gallons of pool
water. So you will add a very small amount, like teaspoons or tablespoons
of it. It will kill the algae, if that's really the problem. Any pool
owner can help you that uses iso-cyanurates for hypochlorous acid
(chlorine) disinfection. He might give you a few mls and save you the trip
to the pool store for a 40 year chiller supply of QAS long chained
surfactant.

You can state, "I have plastic pipes, so where is the copper coming from in
my system?" Our chiller coils were made of copper tubing and the reservoir
tank had copper tubing showing in our units. They're the source of the
copper ions. In our system we also had copper supply lines going to
various labs. A mistake! There are other copper sources too. Ask your
serviceman about potential EM sources like embedded lens cooling lines of
fittings.
Air is normally the source of the CO2 because the tank is normally open to
air.

} From experience, we learned the hard way to run distilled water. I then
put a tablespoon of bicarbonate of soda in the water in case the chiller
water tried to go acidic. This trace ion concentration cuts down on the
corrosive nature of pure distilled water. This bicarb is not necessary,
unless you know the pH in your system drops over time. "Isn't that the
source of the carbonate?" No, we had the problem when we used chloramine-T
and no bicarb. We used bicarb after we switched to distilled water for
trace ions and to slow down any decrease in pH.

You can get into a green color spiral with chloramine-T (CT). You add CT
and the water eventually turns green. You measure the available chlorine
after awhile and it's too low. You add more CT and the available chlorine
goes up. After some time the available chlorine drops and the g