From: Richard Edelmann :      edelmare-at-MUOhio.edu
Date: Tue, 1 Mar 2005 08:12:22 -0500
Subject: [Microscopy] Re: TEM section stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ursula:


Several years ago I compared potassium permanganate post-section
staingin with barium post-section staining, I found barium staing to be much
cleaner to use and did an excellent job (superior to potassium permanganate
in my hands) on plant and fungal walls and protein/carbohydate matrices. I
can't put my hands on the procedure immediately but here the reference:


Hoch, H. C. 1977. Use of permanganate of increase the electron opacity

of fungal walls. Mycologia 69:1209-2113.


**************************************************************************

Harvey C. Hoch

Department of Plant Pathology

Barton Laboratory

Cornell University, NYSAES

Geneva, NY 14456-0462




{color} {param} 0100,0100,0100 {/param} { SEQ CHAPTER \h \r 1} {/color}

} Dear all,

{color} {param} 7F00,0000,0000 {/param} }

} Has anyone ever used Potassium Permanganate as a section stain? Apparently it

} has been shown to improve the contrast in certain types of cell in the artery

} wall and I wondered what might be going on in these cells that their contrast

} should be so improved with this stain. Has anyone a method for the staining?

}

} Many thanks

} Ursula

} ----------------

}

} Ursula J. Potter

} Centre for Electron Optical Studies (CEOS)

} Building 3 West 2.15

} The University of Bath

} Claverton Down

} Bath BA2 7AY

} UK

} Tel: 01225 385651

} Email: U.J.Potter-at-bath.ac.uk

}



{nofill}
Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Director
350 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513.529.5712 Fax: 513.529.4243
E-mail: edelmare-at-muohio.edu
http://www.emf.muohio.edu

"RAM disk is NOT an installation procedure."


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 08:20:14 2005



From: Elaine F. Schumacher :      eschumacher-at-mccrone.com
Date: Tue, 1 Mar 2005 08:26:41 -0600
Subject: [Microscopy] Asbestos analysis info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

As a follow-up to Barbara's reply, contact info for Thom Kubik is:

Thomas A. Kubik & Assoc. (TAKA)
P.O. Box 208
Greenlawn, NY 11704
Phone: 516-261-2117

Another contact is Peter M. Cooke:

Microscopy Instruction, Consultation & Analysis (MICA)
5807 N. Maplewood
Chicago, IL 60659
Phone: 773-334-2240
pmcooke-at-xsite.net

Regards,

Elaine

*********************************************************************
Elaine F.Schumacher
Senior Research Scientist
McCrone Associates, Inc.
850 Pasquinelli Drive
Westmont, IL 60559-5539 USA
630-887-7100 (tel)
630-887-7417 (fax)
E-mail: eschumacher-at-mccrone.com
Web Site: www.mccrone.com
*********************************************************************
This message and any attachments are solely for the
intended recipient. If you are not the intended recipient,
disclosure, copying, use or distribution of the information
included in this message is prohibited.
*********************************************************************

-----Original Message-----
} From: sghoshro-at-NMSU.Edu [mailto:sghoshro-at-NMSU.Edu]
Sent: Monday, February 28, 2005 4:16 PM
To: MSA Listserve

Dear Everyone,

I am interested to know how and where to obtain training to perform TEM,

SEM, and light microscopic analysis of asbestos. Are there short courses

or certifications available to perform asbestos analysis ? Do we need
any
acreditation to perform this sort of analysis ? If any one knows of good

references on asbestos analysis, books, protocols etc. then please let
us
know. We will really appreciate your help.

Thanks in advance.

Soumitra

*************************************************************
Soumitra Ghoshroy
College Associate Professor, Biology
Director, Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
http://biology-web.nmsu.edu/ghoshroy/ghoshroy.htm
http://emldata.nmsu.edu/eml/




From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 09:16:42 2005



From: Randy Boltin :      rboltin-at-mvainc.com
Date: Tue, 01 Mar 2005 10:24:55 -0500
Subject: [Microscopy] Re: Asbestos analysis info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Soumitra:

McCrone Research Institute in Chicago (URL: www.mcri.org) and MICA (E-mail:
PMCOOKE-at-EARTHLINK.NET) are good sources for beginning courses in asbestos analysis
by polarized light microscopy (PLM) for bulk samples and by phase contrast
microscopy (PCM) for air samples by the NIOSH 7400 method. MVA Scientific
Consultants (the company I work with) offers a "Fundamentals of Asbestos Analysis
by Transmission Electron Microscopy" course at our facility in the Atlanta area.
You can check our website (www.mvainc.com) or call us at 770-662-8509 for more
information.

As for accreditation, the National Voluntary Laboratory Accreditation Program
(NVLAP) administered by the National Institute of Standards and Technology (NIST)
offers program-specific accreditations for asbestos analysis by PLM and by TEM
using the approved EPA methods. The NVLAP program for asbestos was mandated by the
Asbestos Hazards Emergency Response Act of 1986 for laboratories conducting
analysis for asbestos materials in school buildings. Lab accreditation is also
available through the American Association for Laboratory Accreditation (A2LA).
Both NVLAP and A2LA provide accreditation according to ISO17025 criteria. You may
want to note that much contract work these days requires that a lab be accredited.

Good reference material on the principles of bulk asbestos analysis by PLM is
available through MRI. EPA Test Method EPA/600/R-93/116: Method for the
Determination in Bulk Building Materials, is available from EPA. The TEM protocol
for analysis of asbestos in air samples is located in 40 CFR Part 763
(Asbestos-Containing Materials in Schools; Final Rule and Notice) Appendix A to
Subpart E.

I hope this helps. Good luck.

Best Regards,
Randy Boltin

sghoshro-at-NMSU.Edu wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Dear Everyone,
}
} I am interested to know how and where to obtain training to perform TEM,
} SEM, and light microscopic analysis of asbestos. Are there short courses
} or certifications available to perform asbestos analysis ? Do we need any
} acreditation to perform this sort of analysis ? If any one knows of good
} references on asbestos analysis, books, protocols etc. then please let us
} know. We will really appreciate your help.
}
} Thanks in advance.
}
} Soumitra
}
} *************************************************************
} Soumitra Ghoshroy
} College Associate Professor, Biology
} Director, Electron Microscopy Lab
} Box 3EML
} New Mexico State University
} Las Cruces, NM 88003
} Tel: 505-646-3268 (office), 646-3283 (lab)
} Fax: 505-646-3282
} e-mail:sghoshro-at-nmsu.edu
} http://biology-web.nmsu.edu/ghoshroy/ghoshroy.htm
} http://emldata.nmsu.edu/eml/

--
-----------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------

The information in this email is confidential and may be legally privileged.
It is intended solely for the addressee. If you are not the intended recipient,
please delete the email and notify MVA Scientific Consultants of the transmission
error.
MVA Scientific Consultants - 5500 Oakbrook Pkwy Suite 200, Norcross, GA 30093 -
(770)662-8509
-------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------





From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 09:54:54 2005



From: David Patton :      David.Patton-at-uwe.ac.uk
Date: Tue, 01 Mar 2005 16:01:13 +0000
Subject: [Microscopy] TEM section stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

When I started here in 1989 potassium permanganate was employed instead
of UA for safety reasons. I formed the impression from the literature
that UA would suit us better and discontinued use of permanganate.

We used it as either:
1% (aq) or 1% in phosphate buffer pH 6.5

It is supposed to be good for membranes but I did not do a rigourous
comparison.

Hayat (1989) (3rd Ed.) Principles and Techniques of Electron Microscopy.
Biological Applications covers it (pp280-281). Hayat describes using it
as part of a triple stain with permanganate and lead citrate.

Dave


-----Original Message-----
} From: Ursula Potter [mailto:U.J.Potter-at-bath.ac.uk]
Sent: 28 February 2005 17:27
To: Microscopy-at-microscopy.com

Dear all,

Has anyone ever used Potassium Permanganate as a section stain?
Apparently
it has been shown to improve the contrast in certain types of cell in
the
artery wall and I wondered what might be going on in these cells that
their
contrast should be so improved with this stain. Has anyone a method for
the
staining?

Many thanks
Ursula
----------------

Ursula J. Potter
Centre for Electron Optical Studies (CEOS)
Building 3 West 2.15
The University of Bath
Claverton Down
Bath BA2 7AY
UK
Tel: 01225 385651
Email: U.J.Potter-at-bath.ac.uk



This incoming email to UWE has been independently scanned for viruses
and any virus detected has been removed using McAfee anti-virus software


This email has been independently scanned for viruses and any virus software has been removed using McAfee anti-virus software



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 09:57:37 2005



From: James Pawley :      jbpawley-at-wisc.edu
Date: Tue, 01 Mar 2005 10:04:29 -0600
Subject: [Microscopy] Re: LED illumination for photomicroscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hi all,

I hope that it is not too commercial to tell you that, in the next
edition of The Handbook, there will be a fairly complete discussion
of how LEDs might be used in microscopy in the chapter on Non-laser
Light Sources by Andreas Nolte from Zeiss. At least it shows that the
idea has promise.

And on the cheap CCD front, I have had fair success hooking the
iSight camera from apple (~$140, 640x480, Firewire) to a Zeiss scope
by the simple expedient of removing the rubber "eye-glasses"
protector gasket on the normal 10x high-eyepoint occular and holding
the camera right in front of it with a plastic collar. It covers a
little more than the whole field of view (some black corners).
Although I haven't used it, I am told that one can use $30
"surveillance" software to record time-lapse sequences on a Mac using
this camera. Might be neat for non-microscope uses too.

Cheers,

Jim P.
--
**********************************************
Prof. James B. Pawley, Ph. 608-263-3147
Room 223, Zoology Research Building,
FAX 608-265-5315
1117 Johnson Ave., Madison, WI, 53706
JBPAWLEY-at-WISC.EDU
3D Microscopy of Living Cells Course, June 11-23, 2005, UBC, Vancouver Canada
Info: http:// www.3dcourse.ubc.ca/ Applications due by March 15, 2005


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 10:35:23 2005



From: Aleksandr Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Tue, 01 Mar 2005 16:45:14 +0000
Subject: [Microscopy] TEM: EM Atlas of tissues and organs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear collegues,

I need EM atlas of mammalian tissues and organs. What can you recommend?

--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 11:51:33 2005



From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Tue, 01 Mar 2005 12:58:26 -0500
Subject: [Microscopy] Re: cheap CCD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

A few months ago we posted, in response to a similar question, a link to
our CheapScopeTM at http://www.aecom.yu.edu/aif/gallery/cheapscope/index.htm

Here's an update:
http://cammer.net/blog/snowflake03.htm

-Michael


} And on the cheap CCD front, I have had fair success hooking the iSight
} camera from apple (~$140, 640x480, Firewire) to a Zeiss scope by the
} simple expedient of removing the rubber "eye-glasses" protector gasket on
} the normal 10x high-eyepoint occular and holding the camera right in
} front of it with a plastic collar.

____________________________________________________________________________
Michael Cammer Analytical Imaging Facility Albert Einstein Coll. of Med.
Jack & Pearl Resnick Campus 1300 Morris Park Ave. Bronx, NY 10461
(718) 430-2890 Fax: 430-8996 URL: http://www.aecom.yu.edu/aif/
**This electronic transmission contains information that is privileged.**



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 11:53:34 2005



From: Mary Albrecht :      malbrecht-at-SSCI-INC.com
Date: Tue, 1 Mar 2005 13:00:47 -0500
Subject: [Microscopy] cGMP microscopy labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello

Are there any contract labs out there that analyse under cGMP conditions?
I'm specifically looking for analysing some pharmaceutical powders by SEM.

Thanks

Mary Albrecht
SSCI, Inc.


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 12:58:09 2005



From: john hoffpauir :      hoffpajo-at-yahoo.com
Date: Tue, 1 Mar 2005 11:04:46 -0800 (PST)
Subject: [Microscopy] Re: TEM: EM Atlas of tissues and organs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

fawcett had one years ago, not certain it is in print,
but should be in your library.
john

--- Aleksandr Mironov
{Aleksandr.Mironov-at-manchester.ac.uk} wrote:

}
}
}
------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The
} Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help
}
http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
-------------------------------------------------------------------------------
}
} Dear collegues,
}
} I need EM atlas of mammalian tissues and organs.
} What can you recommend?
}
} --
} Aleksandr Mironov
} Experimental Officer
} G450A, Stopford Building
} EM Unit, Faculty of Life Sciences
} University of Manchester
} Oxford Road
} Manchester
} M13 9PT
} UK
}
} Tel. +44-(0)161-275-7395
} Fax. +44-(0)161-275-5171
} E-mail: Aleksandr.Mironov-at-manchester.ac.uk
} MSN: amironov-at-hotmail.co.uk
} Web: http://www.empgu.man.ac.uk
}
}
}
}


__________________________________________________
Do You Yahoo!?
Tired of spam? Yahoo! Mail has the best spam protection around
http://mail.yahoo.com


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 13:37:19 2005



From: Tom Phillips :      phillipst-at-missouri.edu
Date: Tue, 01 Mar 2005 13:45:40 -0600
Subject: [Microscopy] Re: TEM: EM Atlas of tissues and organs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Don Fawcett's "The Cell" has the finest TEM views of organelles and
membrane specializations. But is not really an atlas of tissues and
organs. Kessel and Kardon is the best SEM atlas I know.


At 10:45 AM 03/01/05, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thomas E. Phillips, PhD
Professor of Biological Sciences
Director, Molecular Cytology Core
3 Tucker Hall
University of Missouri
Columbia, MO 65211-7400

573-882-4712 (office)
573-882-0123 (fax)
PhillipsT-at-missouri.edu




From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 15:55:07 2005



From: William P. Sharp :      wsharp-at-asu.edu
Date: Tue, 01 Mar 2005 15:01:55 -0700
Subject: [Microscopy] Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy Listers -

Our perennial lament - does anyone out there know of a University or
Institute (non-profit) that operates an EM facility or a Bioimaging
facility that takes in enough money through charge back of users to cover
most or all of the costs of lab operation, to include service contracts on
major instruments? If so, may we know the rates charged? Are there a
relatively large number of outside users that are billed at a much higher
rate than inside users?

We at ASU are in the second year of ramping up the rates charged for zero
to an average charge rate over four years and are feeling some pressure to
charge enough to pay for all operations of our bioimaging facilities, EM
and light as soon as possible. Those of us who are responsible for the
facilities are concerned about pricing ourselves out of the market and we
need some idea of whether it can be done without a total upheaval of
service or the loss of our core facilities.

Thanks for any information or advice you may have for us -

Bill Sharp

William P. Sharp
Arizona State University
School of Life Sciences, box 4501
Tempe, AZ 85287-4501
Phone - (480)-965-3210
Fax - (480)-965-6899



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 16:02:17 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Tue, 01 Mar 2005 17:08:40 -0500
Subject: [Microscopy] Re: TEM: EM Atlas of tissues and organs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

My all-time favorite for this is Johannes A G Rhodin's "Histology:A
Text and Atlas" copyright 1974 Oxford Univ. Press.
It may be out of print, but your library should have it. You may be
able to scare one up on Amazon.
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 20:13:59 2005



From: Anthony Garratt-Reed :      tonygr-at-MIT.EDU
Date: Tue, 01 Mar 2005 21:19:31 -0500
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Bill-

We cover 82%. We are principally a Materials facility, and are a lot more
than imaging (you can see what we have at http://prism.mit.edu).

The trick is that you have to pare down on the service you provide your
users, and you can't be paranoid about them making mistakes. Virtually all
use is "self use" where the users are operating the instruments.

As one example, we operate 4 200KV TEM's (one with EDX), a VG STEM, and 3
SEM's of varying capability (all with EDX, one with EBSD and ESEM),
together with 2 ion mills, a cryomicrotome and a host of other ancilliary
equipment, with 2.4 FTE's. All EM's except the VG are under manufacturer's
contract.

The staff give users very basic operating instructions, then let them
go. We spend most of our time fixing mistakes, telling users that the
instrument won't work because they haven't turned it on, and in similar
pursuits, and we rely on our service contracts to bring in technicians to
fix broken things. You'd be surprised how often we have to explain that an
image is not in focus if it doesn't look sharp!

There are downsides to this philosophy, of course. However, it has a
number of major benefits: the users are forced to learn (and hopefully
understand) the techniques they are employing (after all, we are a
University!); surprisingly, productivity seems to be much higher than when
we were more circumspect; and, of course, we can operate within a
manageable budget (our fiscal director would like the subsidy to be zero,
but we seem to have come to a workable equilibrium - at least for the
present - with the 18%).

Users do break things, but we find that we lose more productivity by
limiting their access than we do by fixing the mistakes. Since our raison
d'etre is to generate results, we opt for the productivity.

One perspective. I hope it helps.

Tony Garratt-Reed.


At 03:01 PM 3/1/2005 -0700, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

*********************************************
Anthony J. Garratt-Reed, M.A., D.Phil
MIT Room 13-1027
77 Massachusetts Avenue
Cambridge, MA 02139-4307
*********************************************
Phone: (617) 253-4622
Fax: (617) 258-6479
*********************************************



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 1 20:48:10 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Tue, 01 Mar 2005 18:55:52 -0800
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

William
In my opinion, it's impossible to make EM facility profitable or
self-sustained. It was discussed at ListServer many times and the
conclusion was that most EM facilities need external funding to survive.
What we usually recover from the users is 30-50% at the best. At least,
this is my experience and opinion. Sergey

At 02:01 PM 3/1/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 04:56:13 2005



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Wed, 2 Mar 2005 11:59:45 +0100
Subject: [Microscopy] Goldstein "Scanning elctron microscopy" Quest.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all

Having the 1980th edition of Godstein et al. Scanning electron microscopy
book, I wanted to buy the new one, and I am waiting since mid November.
The bookseller says me the book would be in reprint. But on the other
side, people from Springeronline saiy they have in on stock (first print
?)... With the confidence we can have on web site informations (amzon or
editors).

I asked me if someone has soon seen the 2005 reprint of that book, or if
it is still being printed.

Thanks

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 05:11:30 2005



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Wed, 2 Mar 2005 12:14:30 +0100
Subject: [Microscopy] Re: LED illumination for photomicroscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Have a look at

http://www.microscopy-uk.org.uk/mag/artaug03/jwled.html

Some informations about white LED illumination. From my experience, it's
not easy to focus a lot of LEDs on the same place. But it's very
interseting for field instruments, working with NiCd cells.


J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr

On Mon, 28 Feb 2005, Roger Baker wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} I recently purchased a Chinese trinocular biological microscope for
} general use. It was an overall good deal with practical design, good
} optics, and flat fields and sharp focus out to the edge, although they
} cut some corners on sturdyness of construction.
}
} I also bought an inexpensive ($ 300) Chinese microscopy camera called
} the Moticam 1000 with which I am not so happy because of a low
} sensitivity CD chip and the fact that there is solarizing at high
} illumination levels and the image pixelates when you use the full one
} mega-pixel resolution. But I try to live with that. Considerably better
} pictures are made by simply holding my 4 megapixel Canon Powershot A80
} up to the eyepiece and shooting.
}
} The high numerical aperture objectives means that the depth of focus is
} small; this scope is solely designed to study flat transparent sections
} mounted on slides. The built-in halogen illumination presents few
} design problems because few photons are lost along the light path. One
} bright white LED inserted in the same light path can nicely substitute
} for their halogen bulb illuminator.
}
} But I want to do reflectance illumination photography of flat-ground
} sections of carbonate micro-fossils stained with gentian violet. Then I
} put on a drop of glycerine and top with a coverslip. This makes the
} denser structure of micro fossil details show up nicely against a deep
} purple matrix, but it takes strong reflective illumination, especially
} with my rather insensitive CCD camera.
}
} I think the standard solution is an illuminator with a halogen light in
} a box and two fiber optic bundles on stalks that can be arranged to
} apply light to two sides of the field. This however costs as about as
} much as my $750 microscope, so I have been investigating LEDS as an
} inexpensive alternative.
}
} What I have found is that you can use a ring of four white LEDs
} (superbrightleds.com in Saint Louis) ) mounted on a fixture that clamps
} on the objective. This works pretty nicely for visual work, since the
} eye is quite sensitive at low light levels. And you can easily power
} this gadget with 5 volts DC; each LED should its own 100 ohm current
} limiting resistor.
}
} But this is still not bright enough for my CCD cameras.
}
} The best fix I have found for the sensitivity problem is to use a ring
} of bright green LEDs.
} I have four mounted now, but a ring of seven would still fit and would
} be better. Each has to be independently adjusted by bending its leads
} to cast its light onto the same small area under the objective. Since
} the purple stain gives a monochromatic image anyhow, the resulting
} image still shows up nicely under both visual examination and for
} photography by my hand held digital camera, and even marginally for my
} Moticam 1000.
}
} Green LEDs are a definite improvement over white LEDs where my CCDs are
} concerned.
} Has LED illumination been discussed on this list before?
}
} Are there any microscope cameras that can compete in both cost and
} image quality with a consumer camera held up to the eyepiece? -- Roger,
} Austin Tx
}
}




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 06:33:22 2005



From: Basil Julies :      bjulies-at-uwc.ac.za
Date: Wed, 02 Mar 2005 14:42:17 +0200
Subject: [Microscopy] Hitachi 650 SEM and 800 TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi everybody

Do you know of anyone who has a broken Hitachi 650 SEM or 800 TEM? I'm
contacting you all the way from South Africa. We cannot afford new
instruments and desperately need to keep the existing ones alive by
using spare parts from other similar obsolete instruments. These
instruments are more than 20 years old and some of the electronic
components are no longer available.
This is a desperate plea for help. Even though money is very scarce on
our side, we are nevertheless willing to pay something towards these
spare parts.

Urgently awaiting a reply

Yours sincerely

Basil Julies





Dr. Basil Julies
Director
Electron Microscope Unit
Physics Department
University of the Western Cape
Private Bag X17
Bellville 7535 South Africa
Tel : (27)(21) 959 2327 or 959 3458
Fax : (27)(21) 959 1335 or 959 3474


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 09:45:42 2005



From: Anita Garg :      Anita.Garg-at-grc.nasa.gov
Date: Wed, 02 Mar 2005 10:52:23 -0500
Subject: [Microscopy] Etching NiTiPt alloys

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear All
Does anybody have experience in etching NiTiPt alloys (~ 25at.%Pt)to reveal
martensite and/or grain boundaries? Any help would be highly appreciated.
TIA
Anita



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 10:14:43 2005



From: Evgenia Pekarskaya :      pekarskaya-at-mail.pse.umass.edu
Date: Wed, 2 Mar 2005 11:21:40 -0500
Subject: [Microscopy] Goldstein "Scanning elctron microscopy" Quest.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Jacques,

I have the third edition of "Scanning Electron Microscopy and X-ray
Microanalysis" by Goldstein et al. published in 2003 by Kluwer Academic
(ISBN: 0-306-47292-9). I believe that this is the latest edition.

Evgenia Pekarskaya

========================
Evgenia Pekarskaya
Keck Electron Microscopy Laboratory
Polymer Science & Engineering
UMass, Amherst; Conte A229
Tel. 413 545 2261
E-fax 325 202 7338
http://www.umassmicroscopy.com/
========================

-----Original Message-----
} From: Faerber Jacques [mailto:Jacques.Faerber-at-ipcms.u-strasbg.fr]
Sent: Wednesday, March 02, 2005 6:00 AM
To: Microscopy Society of America

Hi all

Having the 1980th edition of Godstein et al. Scanning electron microscopy
book, I wanted to buy the new one, and I am waiting since mid November.
The bookseller says me the book would be in reprint. But on the other
side, people from Springeronline saiy they have in on stock (first print
?)... With the confidence we can have on web site informations (amzon or
editors).

I asked me if someone has soon seen the 2005 reprint of that book, or if
it is still being printed.

Thanks

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr








From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 10:37:56 2005



From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Wed, 02 Mar 2005 11:44:45 -0500
Subject: [Microscopy] re: cheap ccd

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

For those of you who aren't completely sick of this discussion...

SPIE oemagazine March 2005 p. 28 is an article by Lionel Baker
lionelbaker-at-ntlworld.com who gives technical details on mounting one ocular
of binoculars to a consumer digital camera to produce 24X telescopic
optical zoom.

____________________________________________________________________________
Michael Cammer Analytical Imaging Facility Albert Einstein Coll. of Med.
Jack & Pearl Resnick Campus 1300 Morris Park Ave. Bronx, NY 10461
(718) 430-2890 Fax: 430-8996 URL: http://www.aecom.yu.edu/aif/
**This electronic transmission contains information that is privileged.**



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 10:50:21 2005



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Wed, 02 Mar 2005 11:51:18 -0500
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Bill,

We are all struggling with the same problem!
We have had to raise charges here too, yet are
only recovering about 60-70% of our cost at the
moment. And that does not include some salaries
that are covered by external grants. If these are
put into the equation, then we are recovering less
than 50%.
Doubling the charges is not an option, as we are
already losing some users due to our current charges.
Only 3 solutions that I can see:
- Boost productivity, but that requires some investment
in new, more sophisticated instrumentation (such as
new CCD camera for the TEM, etc);
- Rely more on NIH, PPG grants, etc, to cover facility
staff salaries;
- Some forms of subsidies from the university. An
important argument is that a good imaging facility
greatly enhances the ability of a university to attract
top faculty, and help faculty get more external
funding, a good portion of which goes back to the
university in the form of overheads! So investing in
a facility should make sense, although I realize that
your typical administrative office might be quite
unmoved by that king of logic!

Good luck

Marc


On Tuesday, March 1, 2005, at 05:01 PM, William P. Sharp wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Microscopy Listers -
}
} Our perennial lament - does anyone out there know of a University or
} Institute (non-profit) that operates an EM facility or a Bioimaging
} facility that takes in enough money through charge back of users to
} cover most or all of the costs of lab operation, to include service
} contracts on major instruments? If so, may we know the rates charged?
} Are there a relatively large number of outside users that are billed
} at a much higher rate than inside users?
}
} We at ASU are in the second year of ramping up the rates charged for
} zero to an average charge rate over four years and are feeling some
} pressure to charge enough to pay for all operations of our bioimaging
} facilities, EM and light as soon as possible. Those of us who are
} responsible for the facilities are concerned about pricing ourselves
} out of the market and we need some idea of whether it can be done
} without a total upheaval of service or the loss of our core } facilities.
}
} Thanks for any information or advice you may have for us -
}
} Bill Sharp
}
} William P. Sharp
} Arizona State University
} School of Life Sciences, box 4501
} Tempe, AZ 85287-4501
} Phone - (480)-965-3210
} Fax - (480)-965-6899
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 14:24:50 2005



From: sghoshro-at-NMSU.Edu
Date: Wed, 2 Mar 2005 13:31:45 -0700 (MST)
Subject: [Microscopy] Thank you for asbestos info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thanks to all of you who responded to my asbestos query. It was indeed a
big help and very informative. Thanks for taking the time to respond and
we will get in touch with some people as suggested by some of you.

This group is just wonderful.

Best wishes,

Soumitra


*************************************************************
Soumitra Ghoshroy
College Associate Professor, Biology
Director, Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
http://biology-web.nmsu.edu/ghoshroy/ghoshroy.htm
http://emldata.nmsu.edu/eml/


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 14:48:22 2005



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Wed, 02 Mar 2005 15:54:59 -0500
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Bill,
We have run into the same problem. After years of having EVERYTHING
covered we are under increased pressure to cover everything. This has
gotten worse as University budgets have shrunk. We cover service contracts
for 3 microscopes (~$60,000) and additional for a full service and
multi-user facility with TEM, SEM, and LM. We still fall about 15-20% short
of covering everything (including data and telephone lines, postage, all
consumables, etc).

The administration's response is to increase rates. I expect increasing
rates will result in decreased user hours and no gain in revenue...until you
get to the point where users drop off and revenue decreases. Unfortunately
non-microscopists do not understand that this is not a speed discipline. It
takes time to adequately align a scope and thoroughly examine samples. If a
user is concerned about watching the clock than quality of research is sure
to suffer. Aren't the facilities here to assist in generating good
research? Or maybe I have been mistaken about this all these years. With
hundreds of thousands of dollars invested in instrumentation, you would
think that universities could find 10-20 thousand each year to help keep
them accessible to the maximum number of researchers.

Debby

P.S. I am saving responses to Bill's E-mail and intend to give them all to
my administrators and advisory committee. I think they can do a lot to help
support my arguments so keep them coming!

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy


On 3/1/05 5:01 PM, "William P. Sharp" {wsharp-at-asu.edu} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Microscopy Listers -
}
} Our perennial lament - does anyone out there know of a University or
} Institute (non-profit) that operates an EM facility or a Bioimaging
} facility that takes in enough money through charge back of users to cover
} most or all of the costs of lab operation, to include service contracts on
} major instruments? If so, may we know the rates charged? Are there a
} relatively large number of outside users that are billed at a much higher
} rate than inside users?
}
} We at ASU are in the second year of ramping up the rates charged for zero
} to an average charge rate over four years and are feeling some pressure to
} charge enough to pay for all operations of our bioimaging facilities, EM
} and light as soon as possible. Those of us who are responsible for the
} facilities are concerned about pricing ourselves out of the market and we
} need some idea of whether it can be done without a total upheaval of
} service or the loss of our core facilities.
}
} Thanks for any information or advice you may have for us -
}
} Bill Sharp
}
} William P. Sharp
} Arizona State University
} School of Life Sciences, box 4501
} Tempe, AZ 85287-4501
} Phone - (480)-965-3210
} Fax - (480)-965-6899
}
}




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 16:47:42 2005



From: John Knowles :      john-at-microvisionlabs.com
Date: Wed, 2 Mar 2005 17:49:38 -0500
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Bill et al,

I own and operate a small analytical microscopy lab and like everyone
else in the dreaded private sector, I have to make a profit every month
or else I do not get paid. Because of this my focus is naturally on
serving the client's microscopy needs and not so much on research.

I was always under the impression that Universities were mainly focused
on research. When part of my tax dollars goes to pay for new equipment
for University microscopy labs I am ok with it as long as I believe that
it is going to be used for education and/or research projects that the
University is doing. When I hear that this equipment is being used to
set up microscopy service labs that are competing against me then I am
definitely not ok with it. Why should I have to pay for all of my
equipment, come in on weekends and stay late into the night doing my own
equipment maintenance and service, and pay 100% of my costs when the
government is using my tax money to help my competition?

There are a number of local Universities in the Boston area that are
happy to accept government grants for new equipment and then use it to
take business away from privately owned labs like mine. I have talked
to many companies that have told me that they go to a University
Microscopy lab to get there work done instead of a lab like mine because
the Universities rates are so low. If I only had to cover 80% (or less)
of my costs and got grants for my equipment I could drop my rates as
well. MIT is one of the few Universities that play by the rules. They
do not take business away from privately own labs like mine. In fact
they send private companies away and recommend labs like mine to them.
And for that I am grateful.

I do not think Universities should compete against private labs when
they are getting government money (just in case you have not figured
that out by now 8^). If they must then at least raise your rates to the
typical rates in your area. That would make it a fair playing field.

Another thought I had, do your administrators factor in any of the
students tuition when calculating your profitability? There are many
students that pick which college they are going to attend based on, at
least in part, what the facilities are at that location. A microscopy
department is a big asset to a school that is trying to attract students
interested in the sciences (e.g. biology and materials). When you look
at a University Microscopy Dept. as a part of the whole education system
then it is silly to think it has to act like a business. The campus
grounds need to be kept up as well in order for a university to attract
and keep students. Does your administration ask the grounds crew to be
profitable as well? Maybe compete with locate landscaping companies
mowing lawns and trimming bushes. I bet not.

John Knowles
President
MicroVision Laboratories, Inc.


-----Original Message-----
} From: Debby Sherman [mailto:dsherman-at-purdue.edu]
Sent: Wednesday, March 02, 2005 3:55 PM
To: William P. Sharp; Microscopy-at-microscopy.com

Bill,
We have run into the same problem. After years of having EVERYTHING
covered we are under increased pressure to cover everything. This has
gotten worse as University budgets have shrunk. We cover service
contracts
for 3 microscopes (~$60,000) and additional for a full service and
multi-user facility with TEM, SEM, and LM. We still fall about 15-20%
short
of covering everything (including data and telephone lines, postage, all
consumables, etc).

The administration's response is to increase rates. I expect
increasing
rates will result in decreased user hours and no gain in revenue...until
you
get to the point where users drop off and revenue decreases.
Unfortunately
non-microscopists do not understand that this is not a speed discipline.
It
takes time to adequately align a scope and thoroughly examine samples.
If a
user is concerned about watching the clock than quality of research is
sure
to suffer. Aren't the facilities here to assist in generating good
research? Or maybe I have been mistaken about this all these years.
With
hundreds of thousands of dollars invested in instrumentation, you would
think that universities could find 10-20 thousand each year to help keep
them accessible to the maximum number of researchers.

Debby

P.S. I am saving responses to Bill's E-mail and intend to give them all
to
my administrators and advisory committee. I think they can do a lot to
help
support my arguments so keep them coming!

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy


On 3/1/05 5:01 PM, "William P. Sharp" {wsharp-at-asu.edu} wrote:

}
}
}
------------------------------------------------------------------------
------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe --
http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
------}
-
}
} Microscopy Listers -
}
} Our perennial lament - does anyone out there know of a University or
} Institute (non-profit) that operates an EM facility or a Bioimaging
} facility that takes in enough money through charge back of users to
cover
} most or all of the costs of lab operation, to include service
contracts on
} major instruments? If so, may we know the rates charged? Are there a
} relatively large number of outside users that are billed at a much
higher
} rate than inside users?
}
} We at ASU are in the second year of ramping up the rates charged for
zero
} to an average charge rate over four years and are feeling some
pressure to
} charge enough to pay for all operations of our bioimaging facilities,
EM
} and light as soon as possible. Those of us who are responsible for the
} facilities are concerned about pricing ourselves out of the market and
we
} need some idea of whether it can be done without a total upheaval of
} service or the loss of our core facilities.
}
} Thanks for any information or advice you may have for us -
}
} Bill Sharp
}
} William P. Sharp
} Arizona State University
} School of Life Sciences, box 4501
} Tempe, AZ 85287-4501
} Phone - (480)-965-3210
} Fax - (480)-965-6899
}
}




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 17:10:34 2005



From: George Langford, Sc.D. :      amenex-at-amenex.com
Date: Wed, 2 Mar 2005 18:17:33 -0500 (EST)
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Microscopists !

It used to be that my hackles would rise whenever this sort of topic
started. However, in the present instance the discussion has been
healthy, albeit depressing. When this starts at the research facilities
of major companies, one knows that big layoffs are certain to follow,
because the lab is now considered expendable ...

While my company (Amenex Associates) was going strong, three of
us managed to keep an ETEC Autoprobe busy, along with associated
metallographic imaging and specimen preparation facilities, located
in about 3000 square feet of office space. We had enough consulting
business to make living wages for about fifteen years. Then the
landlord wanted more money, a longer lease, and separate utilities
billing. That was it. Now my business partner and I work out of our
homes. We send our lab work to a small private lab which seems also
to be hanging on with even a little left over to buy new equipment.

The real problem is with rising inefficiency at the management level,
the same force that's causing college tuition to rise astronomically.
I remember the ridiculous overhead rates - even overhead on tuition
payments for my graduate students - such that an umpteen dollar
research contract yielded mebbe a thousand dollars a year for new
equipment or consumables.

George Langford, Sc.D.
Principal Consultant
Amenex Associates, Inc.
amenex-at-amenex.com
http://www.amenex.com/


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 18:27:46 2005



From: Alan Stone :      as-at-astonmet.com
Date: Wed, 02 Mar 2005 18:34:18 -0600
Subject: [Microscopy] Re: RE: Re: Cost Recovery in Imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


John,

I also have my own small consulting lab and face the same issues. There
are others in the group who know about this, but I believe that NSF grants
include a clause taking in outside work not be at the expense of local
commercial facilities. We are a small community and we should expect our
academic peers to honor the terms of their grant money.

I would suggest that you ensure that the local academics are fully aware of
your capabilities so they have the excuse that they are unaware of your
existence is not valid.

Alan Stone
ASTON Metallurgical Services Co., Inc.


At 04:49 PM 3/2/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 18:50:08 2005



From: Bryony James :      b.james-at-auckland.ac.nz
Date: Thu, 03 Mar 2005 13:57:02 +1300
Subject: [Microscopy] Re: Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

This debate seems to crop up with a worrying increase in frequency. The
argument that good research and good researchers require access to
state-of-the-art facilities goes without saying. The only people who
can't seem to see this are University administrators and those who reply
"surely good research brings in money to pay user charges?" (where do
you even start with that question?).

In my case I run a facility that is tiny by US standards but fully
tooled up by the standards of New Zealand (1 FEG-SEM +EDS +cryo, 1
FEG-ESEM +EDS +EBSD, 1 imaging XPS and 1 AFM). I think we are almost
unique in that the University expects us to cover not only salaries,
consumables and maintenance costs but all our depreciation as well.
Each year I have come up with a new financial "model" that includes
direct contributions from the Faculties in order to break even. This
year I have been told that is not acceptable and user charges have to
cover everything. It's going to be an interesting year, anyone want to
buy a kidney?

I would like to ask the listserv has anyone ever done (or seen done) the
financial analyis of what one published paper is worth to a tertiary
institution? Obviously the answer will change with the funding
situation in each country but it would be an interesting comparison.

Cheers
Bryony James

Research Centre for Surface and Materials Science
University of Auckland
New Zealand


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 19:01:49 2005



From: Hicks, Aaron :      A.W.Hicks-at-massey.ac.nz
Date: Thu, 3 Mar 2005 14:08:40 +1300
Subject: [Microscopy] Re: RE: Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all,

I've been following this thread with some interest.

Is the situation similar for other facilities in other countries?

Aaron Hicks
Electron Microscopy Preparation Technician

Comparative Physiology and Anatomy
Institute of Veterinary, Animal, and Biomedical Sciences
Massey University

PN-412
Private Bag 11 222
Palmerston North
New Zealand

Phone +64 06 350 4874


-----Original Message-----
} From: Alan Stone [mailto:as-at-astonmet.com]
Sent: Thursday, 3 March 2005 1:34 p.m.
To: John Knowles
Cc: microscopy-at-microscopy.com


John,

I also have my own small consulting lab and face the same issues. There

are others in the group who know about this, but I believe that NSF
grants
include a clause taking in outside work not be at the expense of local
commercial facilities. We are a small community and we should expect our

academic peers to honor the terms of their grant money.

I would suggest that you ensure that the local academics are fully aware
of
your capabilities so they have the excuse that they are unaware of your
existence is not valid.

Alan Stone
ASTON Metallurgical Services Co., Inc.


At 04:49 PM 3/2/2005, you wrote:


} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America

} or else I do not get paid. Because of this my focus is naturally on
} serving the client's microscopy needs and not so much on research.
}
} I was always under the impression that Universities were mainly focused

} on research. When part of my tax dollars goes to pay for new equipment

} for University microscopy labs I am ok with it as long as I believe
} that it is going to be used for education and/or research projects that

} the University is doing. When I hear that this equipment is being used

} to set up microscopy service labs that are competing against me then I
} am definitely not ok with it. Why should I have to pay for all of my
} equipment, come in on weekends and stay late into the night doing my
} own equipment maintenance and service, and pay 100% of my costs when
} the government is using my tax money to help my competition?
}
} There are a number of local Universities in the Boston area that are
} happy to accept government grants for new equipment and then use it to
} take business away from privately owned labs like mine. I have talked
} to many companies that have told me that they go to a University
} Microscopy lab to get there work done instead of a lab like mine
} because the Universities rates are so low. If I only had to cover 80%
} (or less) of my costs and got grants for my equipment I could drop my
} rates as well. MIT is one of the few Universities that play by the
} rules. They do not take business away from privately own labs like
} mine. In fact they send private companies away and recommend labs like

} mine to them. And for that I am grateful.
}
} I do not think Universities should compete against private labs when
} they are getting government money (just in case you have not figured
} that out by now 8^). If they must then at least raise your rates to
} the typical rates in your area. That would make it a fair playing
} field.
}
} Another thought I had, do your administrators factor in any of the
} students tuition when calculating your profitability? There are many
} students that pick which college they are going to attend based on, at
} least in part, what the facilities are at that location. A microscopy
} department is a big asset to a school that is trying to attract
} students interested in the sciences (e.g. biology and materials). When

} you look at a University Microscopy Dept. as a part of the whole
} education system then it is silly to think it has to act like a
} business. The campus grounds need to be kept up as well in order for a

} university to attract and keep students. Does your administration ask
} the grounds crew to be profitable as well? Maybe compete with locate
} landscaping companies mowing lawns and trimming bushes. I bet not.
}
} John Knowles
} President
} MicroVision Laboratories, Inc.
}
}
} -----Original Message-----
} } From: Debby Sherman [mailto:dsherman-at-purdue.edu]
} Sent: Wednesday, March 02, 2005 3:55 PM
} To: William P. Sharp; Microscopy-at-microscopy.com
} Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities
}
}
}
} -----------------------------------------------------------------------
} -
} ------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America

} This has gotten worse as University budgets have shrunk. We cover
} service contracts for 3 microscopes (~$60,000) and additional for a
} full service and multi-user facility with TEM, SEM, and LM. We still
} fall about 15-20% short
} of covering everything (including data and telephone lines, postage,
all
} consumables, etc).
}
} The administration's response is to increase rates. I expect
} increasing rates will result in decreased user hours and no gain in
} revenue...until you
} get to the point where users drop off and revenue decreases.
} Unfortunately
} non-microscopists do not understand that this is not a speed
discipline.
} It
} takes time to adequately align a scope and thoroughly examine samples.
} If a
} user is concerned about watching the clock than quality of research is
} sure
} to suffer. Aren't the facilities here to assist in generating good
} research? Or maybe I have been mistaken about this all these years.
} With
} hundreds of thousands of dollars invested in instrumentation, you would
} think that universities could find 10-20 thousand each year to help
keep
} them accessible to the maximum number of researchers.
}
} Debby
}
} P.S. I am saving responses to Bill's E-mail and intend to give them
} all to my administrators and advisory committee. I think they can do a

} lot to help
} support my arguments so keep them coming!
}
} Debby Sherman, Manager Phone: 765-494-6666
} Life Science Microscopy Facility FAX: 765-494-5896
} Purdue University E-mail: dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907 http://www3.agriculture.purdue.edu/microscopy
}
}
} On 3/1/05 5:01 PM, "William P. Sharp" {wsharp-at-asu.edu} wrote:
}
} }
} }
} }
} -----------------------------------------------------------------------
} -
} ------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
} -----------------------------------------------------------------------
} -
} ------}
} -
} }
} } Microscopy Listers -
} }
} } Our perennial lament - does anyone out there know of a University or

} } Institute (non-profit) that operates an EM facility or a Bioimaging
} } facility that takes in enough money through charge back of users to
} cover
} } most or all of the costs of lab operation, to include service
} contracts on
} } major instruments? If so, may we know the rates charged? Are there a

} } relatively large number of outside users that are billed at a much
} higher
} } rate than inside users?
} }
} } We at ASU are in the second year of ramping up the rates charged for
} zero
} } to an average charge rate over four years and are feeling some
} pressure to
} } charge enough to pay for all operations of our bioimaging
} } facilities,
} EM
} } and light as soon as possible. Those of us who are responsible for
} } the facilities are concerned about pricing ourselves out of the
} } market and
} we
} } need some idea of whether it can be done without a total upheaval of

} } service or the loss of our core facilities.
} }
} } Thanks for any information or advice you may have for us -
} }
} } Bill Sharp
} }
} } William P. Sharp
} } Arizona State University
} } School of Life Sciences, box 4501
} } Tempe, AZ 85287-4501
} } Phone - (480)-965-3210
} } Fax - (480)-965-6899
} }
} }





From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 19:13:57 2005



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Wed, 02 Mar 2005 19:26:17 -0600
Subject: [Microscopy] Re: Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

debbie

if it will help, i have been filing these notes for the past several
years. it won't be fancy, but i can share them all. also, i believe
phil oshel did something on this as a survey 2-3 years ago.

paul



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 19:38:36 2005



From: cosandey-at-scils.rutgers.edu (by way of MicroscopyListserver)
Date: Wed, 2 Mar 2005 19:45:57 -0600
Subject: [Microscopy] viaWWW: Research Associate Position in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (cosandey-at-scils.rutgers.edu) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, March 2, 2005 at 18:59:31
---------------------------------------------------------------------------

Email: cosandey-at-scils.rutgers.edu
Name: Fred Cosandey

Organization: Rutgers University

Title-Subject: [Microscopy] [Filtered] MListserver: Research Associate Position in TEM

Question: Research Associate Position
Transmission Electron Microscopy

The Department of Ceramic and Materials Engineering at Rutgers University is currently seeking a candidate for a Research Associate position in Transmission Electron Microscopy.
The successful candidate should have a PhD degree in Materials Science or related discipline and have extensive experience in the use of transmission electron microscope including HAADF, GIF and EELS spectroscopy techniques. Prior experience with JEOL 2010F STEM would be preferable.
Specific projects involve EELS analysis of nanostructured electrode materials used in Li-Ion batteries, including EELS determination of valence state and, structural and chemical analysis of electrodes during Li-induced phase transformations. Also, chemical and electronic structure determination of interfaces in oxides and metal-oxides systems using HAADF STEM and EELS.

The appointment will be for one year and may be renewable for subsequent years. Interested candidate should send their CV with three references to:

Professor Frederic Cosandey
Department of Ceramic and Materials Engineering
Rutgers University
607 Taylor Rd.
Piscataway, NJ 08854-8065,
cosandey-at-rscils.rutgers.edu, 732-445-4942 (phone), 732-445-5977 (fax)

Rutgers University is an equal opportunity/affirmative action employer.


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 2 21:41:20 2005



From: Rosemary White :      Rosemary.White-at-csiro.au
Date: Thu, 03 Mar 2005 14:48:57 +1100
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear all,

Ah yes, the same old chestnut keeps coming from administrators.
My response to this was as follows.

Because we're a small unit (TEM, SEM, confocal, other LMs, 2 staff) I can
keep pretty good track of all the work going through. I keep very good
track of the publications that come out using work done in Microscopy.

Now, our Division of CSIRO - Plant Industry - has an income of about $70
million per year (about 60% federal, 40% industry). The cost of maintaining
Microscopy is about $370,000 per year (I'm expected to recover two salaries,
depreciation on equipment - which is about half the total, consumables,
rental on the building, all utilities, etc. - the lot). Of the publications
from the Division of Plant Industry, at least 10% use Microscopy to some
degree. I worked out that, on average, Microscopy's contribution to these
publications was about 2% per year, minimum - between 5% and 80% of each of
those 10% of publications. Now 2% of $70 million is about $1.4 million that
Microscopy might expect as input to support this output - i.e. just under 4
times our actual cost.

QED - we are way overperforming, so how can it be argued that we cost too
much? (Yes, it's pretty rough, and of course, I haven't included the salary
time others have spent in Microscopy using the equipment....) In our case,
it's an accounting exercise - our costs should be associated with a project
number rather being "general costs", but the problem is, if we increase our
charges to internal users, our external, publicly-funded collaborators will
find us way too expensive to work with.

I suspect that the need to account for an expense against some specific
account number may underlie some of these demands for full cost recovery.
Our accountants hate having expenses that they can't allocate to a
particular grant or project number. Because we (Microscopy) recover only
about 50% of costs directly from users, it means that the accountants have
to somehow spread the rest of the costs across all the other numbers they've
got, and they don't like doing that (although they're quite happy to do it
for adminstration costs!). However, since I pointed out our productivity
vs. input as above (and a few other things), I've had no more hassles from
admin....yet.

It may not work in all cases, and for large units may not be possible (too
many users to keep track of), but it might be worth pointing out how much of
the research output that contributes to your institution's reputation goes
through the microscopy unit.

good luck,
Rosemary

Dr. Rosemary White rosemary.white-at-csiro.au
Microscopy Centre ph. 61-2-6246 5475
CSIRO Plant Industry mob. 61-0402 835 973
GPO Box 1600 fax. 61-2-6246 5334
Canberra, ACT 2601
Australia


} From: Debby Sherman {dsherman-at-purdue.edu}
} Date: Wed, 02 Mar 2005 15:54:59 -0500
} To: "William P. Sharp" {wsharp-at-asu.edu} , {Microscopy-at-microscopy.com}
} Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Bill,
} We have run into the same problem. After years of having EVERYTHING
} covered we are under increased pressure to cover everything. This has
} gotten worse as University budgets have shrunk. We cover service contracts
} for 3 microscopes (~$60,000) and additional for a full service and
} multi-user facility with TEM, SEM, and LM. We still fall about 15-20% short
} of covering everything (including data and telephone lines, postage, all
} consumables, etc).
}
} The administration's response is to increase rates. I expect increasing
} rates will result in decreased user hours and no gain in revenue...until you
} get to the point where users drop off and revenue decreases. Unfortunately
} non-microscopists do not understand that this is not a speed discipline. It
} takes time to adequately align a scope and thoroughly examine samples. If a
} user is concerned about watching the clock than quality of research is sure
} to suffer. Aren't the facilities here to assist in generating good
} research? Or maybe I have been mistaken about this all these years. With
} hundreds of thousands of dollars invested in instrumentation, you would
} think that universities could find 10-20 thousand each year to help keep
} them accessible to the maximum number of researchers.
}
} Debby
}
} P.S. I am saving responses to Bill's E-mail and intend to give them all to
} my administrators and advisory committee. I think they can do a lot to help
} support my arguments so keep them coming!
}
} Debby Sherman, Manager Phone: 765-494-6666
} Life Science Microscopy Facility FAX: 765-494-5896
} Purdue University E-mail: dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907
} http://www3.agriculture.purdue.edu/microscopy
}
}
} On 3/1/05 5:01 PM, "William P. Sharp" {wsharp-at-asu.edu} wrote:
}
} }
} }
} }
-----------------------------------------------------------------------------} }
-
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
}
------------------------------------------------------------------------------}
}
} -
} }
} } Microscopy Listers -
} }
} } Our perennial lament - does anyone out there know of a University or
} } Institute (non-profit) that operates an EM facility or a Bioimaging
} } facility that takes in enough money through charge back of users to cover
} } most or all of the costs of lab operation, to include service contracts on
} } major instruments? If so, may we know the rates charged? Are there a
} } relatively large number of outside users that are billed at a much higher
} } rate than inside users?
} }
} } We at ASU are in the second year of ramping up the rates charged for zero
} } to an average charge rate over four years and are feeling some pressure to
} } charge enough to pay for all operations of our bioimaging facilities, EM
} } and light as soon as possible. Those of us who are responsible for the
} } facilities are concerned about pricing ourselves out of the market and we
} } need some idea of whether it can be done without a total upheaval of
} } service or the loss of our core facilities.
} }
} } Thanks for any information or advice you may have for us -
} }
} } Bill Sharp
} }
} } William P. Sharp
} } Arizona State University
} } School of Life Sciences, box 4501
} } Tempe, AZ 85287-4501
} } Phone - (480)-965-3210
} } Fax - (480)-965-6899
} }
} }
}
}
}


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 03:27:43 2005



From: Richard Beanland :      richard.beanland-at-bookham.com
Date: Thu, 3 Mar 2005 09:34:49 -0000
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I just thought I'd say that this kind of problem is not unique to the academic sector, and I have a lot of sympathy for anyone trying to run an independent lab and cover all their costs. When I first started here 10 years ago, we were subject to the same accounting rules as the rest of the business, and had to recover all our costs through user charges - although much of this was in 'wooden dollars'. After a few years, some divisions of the company were spun off into start-ups or moved to other sites, and we started to really struggle to fill in the time sheets every week. We ended up taking problems from engineers inside the company, using our expertise to get some good information, and going back to their manager trying to sell the report. Eventually, after a lot of pushing, we were turned into an overhead for the whole site. The result - a lot more work, more freedom to work on the issues important to key projects and overall a far better contribution to the business. I try to keep a list of the money-burning problems we have helped to solve, just one or two big ones can cover our costs for the year.
The argument that seems to work is "how much money would it cost if we were not here", and although this is intangible, accountants seem much happier with this kind of calculation. Perhaps the same argument can be used in the academic sector - take each paper published which uses microscopy off the list and work out how your institution would then fare in a research assessment exercise, translate that into fewer grants, fewer students, etc. Or the cost of having that work done in a commercial lab. I would think that even being quite cautious with the assumptions you should be able to cover your costs several times over.

Good luck

Richard
________________________________________
Richard Beanland
Analytical Services
Bookham Inc
Caswell
Towcester
Northants
NN12 8EQ
United Kingdom
Tel. +44 1327 356362
Fax. +44 1327 356775
http://www.bookham.com
________________________________________


=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.
No part of this message can be considered a request for goods or
services.
=======================================================================



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 03:57:07 2005



From: Robert H. Olley :      hinmeigeng-at-hotmail.com
Date: Thu, 03 Mar 2005 10:02:59 +0000
Subject: [Microscopy] RE: TEM section stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have played a lot with potassium permanaganate in various conditions, and
most likely when used a a stain one is getting a precipitate of Manganese
Dioxide (Mn02) formed round the most reactice and reducing components of the
cell. I'm not a bio man myself, but I would suggest that the main thing to
play around with is the pH of your staining solution, perhaps using
appropriate buffers. What buffers do you normally use? Citrate, for
example, would be easily oxidised by the permanganate, which would shed Mn02
everywhere.

-----------------------------------
Robert H. Olley
Reply to: R.H.Olley-at-reading.ac.uk
URL: http://www.rdg.ac.uk/~spsolley
-----------------------------------


>Has anyone ever used Potassium Permanganate as a section stain?
>Apparently it has been shown to improve the contrast in certain
>types of cell in the artery wall and I wondered what might be going
>on in these cells that their contrast should be so improved with
>this stain. Has anyone a method for the staining?

>Ursula J. Potter
>Centre for Electron Optical Studies (CEOS)
>Building 3 West 2.15
>The University of Bath
>Claverton Down
>Bath BA2 7AY
>UK
>Tel: 01225 385651
>Email: U.J.Potter-at-bath.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 08:22:13 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Thu, 03 Mar 2005 09:27:24 -0500
Subject: [Microscopy] RE: Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

In response to John's comments about companies using university
facilities because of their low rates...
Our rates are low with respect to commercial firms, BUT that is only
for our "in house" personnel. I am under orders from my dean to bill
ALL outside users (including academics) my usual fees PLUS 70% (the
rate that NIH pays my institution for indirect costs). I believe
that that extra charge brings my rates up to somewhere around what
commercial labs charge. I know of other university facilities that
double their normal rates for all "off campus" users.
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 08:38:32 2005



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Thu, 03 Mar 2005 09:43:30 -0500
Subject: [Microscopy] Re: RE: Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi John,

These are very good points that you are raising.
However, I don't know how widespread a problem
it is. I have only handled projects for outsiders twice
in the last 6 years, and in both cases it was small
local companies that had been set up by ex-
postdocs from my department. I have to admit that
I have been thinking about advertising our services
to the outside as a way to boost our revenues, but
I gave up on the idea not because I wanted to remain
fair to the competition, but mostly because I want
to remain dedicated to the research community in
my university, even at the risk of running deficits. And
to be fair to our administration, they have allowed us
to run deficits without putting too much pressure on
us, and they have never suggested or encouraged us
to seek business from the outside. Actually, I think
they would be against the idea.
Best

Marc

On Wednesday, March 2, 2005, at 05:49 PM, John Knowles wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Bill et al,
}
} I own and operate a small analytical microscopy lab and like everyone
} else in the dreaded private sector, I have to make a profit every month
} or else I do not get paid. Because of this my focus is naturally on
} serving the client's microscopy needs and not so much on research.
}
} I was always under the impression that Universities were mainly focused
} on research. When part of my tax dollars goes to pay for new equipment
} for University microscopy labs I am ok with it as long as I believe
} that
} it is going to be used for education and/or research projects that the
} University is doing. When I hear that this equipment is being used to
} set up microscopy service labs that are competing against me then I am
} definitely not ok with it. Why should I have to pay for all of my
} equipment, come in on weekends and stay late into the night doing my
} own
} equipment maintenance and service, and pay 100% of my costs when the
} government is using my tax money to help my competition?
}
} There are a number of local Universities in the Boston area that are
} happy to accept government grants for new equipment and then use it to
} take business away from privately owned labs like mine. I have talked
} to many companies that have told me that they go to a University
} Microscopy lab to get there work done instead of a lab like mine
} because
} the Universities rates are so low. If I only had to cover 80% (or
} less)
} of my costs and got grants for my equipment I could drop my rates as
} well. MIT is one of the few Universities that play by the rules. They
} do not take business away from privately own labs like mine. In fact
} they send private companies away and recommend labs like mine to them.
} And for that I am grateful.
}
} I do not think Universities should compete against private labs when
} they are getting government money (just in case you have not figured
} that out by now 8^). If they must then at least raise your rates to
} the
} typical rates in your area. That would make it a fair playing field.
}
} Another thought I had, do your administrators factor in any of the
} students tuition when calculating your profitability? There are many
} students that pick which college they are going to attend based on, at
} least in part, what the facilities are at that location. A microscopy
} department is a big asset to a school that is trying to attract
} students
} interested in the sciences (e.g. biology and materials). When you look
} at a University Microscopy Dept. as a part of the whole education
} system
} then it is silly to think it has to act like a business. The campus
} grounds need to be kept up as well in order for a university to attract
} and keep students. Does your administration ask the grounds crew to be
} profitable as well? Maybe compete with locate landscaping companies
} mowing lawns and trimming bushes. I bet not.
}
} John Knowles
} President
} MicroVision Laboratories, Inc.
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 08:54:15 2005



From: Elaine F. Schumacher :      eschumacher-at-mccrone.com
Date: Thu, 3 Mar 2005 09:00:59 -0600
Subject: [Microscopy] Announcing New Short Courses in Light and Electron Microscopy in 2005

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Located in Westmont, Illinois, the McCrone College of Microscopy is an
institution for special instruction and specialized education, whose
goal
is to advance the knowledge and understanding of light and electron
microscopy for materials analysis. It is dedicated to quality hands-on
teaching of micro-analytical principles, techniques and applications,
using state-of-the-art light and electron microscopes and accessories.

The following short courses will be offered in 2005:

MCM200: Scanning Electron Microscopy
Dates: April 4-8, 2005

MCM430: Microscopical Identification of White-Powder Unknowns (Part 1)
Dates: April 18-22, 2005

MCM300: Microscopic Particle Handling: Particle Isolation, Manipulation
and Mounting
Dates: May 9-13, 2005

MCM430: Microscopical Identification of White-Powder Unknowns (Part 1)
Dates: June 6-10, 2005

MCM410: Microscopical Identification of Pharmaceutical Materials and
Contaminants
Dates: July 11-15, 2005

MCM140: Microscopical Particle Analysis
Dates: August 8-12, 2005

MCM400: Microscopical Examination of Forensic Trace Evidence
Particles (Part 1)
Dates: September 12-16, 2005

MCM420: Microscopical Identification of Pigments for Art Conservation
and Architectural Restoration Professionals
Dates: September 19-23, 2005

MCM300: Microscopic Particle Handling: Particle Isolation, Manipulation
and Mounting
Dates: October 3-7, 2005

MCM200: Scanning Electron Microscopy
Dates: October 17-21, 2005

MCM100: Polarized Light and Chemical Microscopy
Dates: November 14-18, 2005

These courses are presented by instructors with distinguished records,
experience and expertise in polarized light microscopy and scanning
electron microscopy. For instructor biographies, course details and
other information, go to www.mccrone.com. Inquiries can be directed to
courses-at-mcccrone.com.

Thank you.

Charles A. Zona
Vice President
Director of Education
McCrone College of Microscopy
McCrone Associates, Inc.
850 Pasquinelli Drive
Westmont, IL 60559
Telephone 630-887-7100
Email: czona-at-mccrone.com
Website: www.mccrone.com




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 10:14:33 2005



From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Thu, 3 Mar 2005 10:21:37 -0600
Subject: [Microscopy] Cost recovery at imaging facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We have also been dealing with this issue for some time, and although we
are under pressure to increase revenues and it is made clear that the
ideal is to recover all costs, the pressure is not unduly severe. Part
of the reason, I think, is that we have been aggressive in recent years
in trying to increase our revenue flow and user base (as I'm sure many
labs have been), and we have quadrupled our usage since 2000.

Currently we recover all but about 30% of direct expenses, as close as I
can tell. This includes service contracts, consumables, and partial
salaries, although not building overhead and equipment replacement,
which of course would take our percentage way down if we included them
as direct costs.

But it is important to note that we have an educational component to our
facility's mission, as well as a research one. We run two courses out
of our lab and assist with lab facilities for an outside class. We
also provide tours, lectures, and workshops, usually at no charge, for
courses, local schools, and other universities. We run fee-based
workshops and short courses on occasion on a break-even basis (if we're
lucky!), and we quite often are used as part of the itinerary for
recruitment of prospective faculty and sometimes graduate students. All
of this represents a lot of time and effort taken away from just serving
our clients' research needs and from new protocol development. How do
we measure the financial value of this to our university?

I think it's safe to say that this issue will never go away, especially
in these precarious budgeting times, but as the replies to this question
are showing, there are many arguments that can be put forward that
research imaging facilities generate great benefits for their
institutions that are not always easy to quantify in dollar terms. That
doesn't mean they don't exist and that doesn't mean we should apologize
for somehow being inadequate in the eyes of the accountants. So far,
I've never been charged a fee for using our university library and I
expect their cost-recovery is....ahem....somewhat lower than ours.

My two coppers worth.

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility---We Do Small Well!
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.emc.missouri.edu








From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 10:58:24 2005



From: Carla M Conway :      cmconway-at-usgs.gov
Date: Thu, 3 Mar 2005 09:04:58 -0800
Subject: [Microscopy] permeabilization needed in IHC of viral protein?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html






Hello,

I am interested in using IHC to detect nucleocapsid (N) protein within
rhabdovirus. I have had success in detecting the glycoprotein on the
envelope surface, however the N protein is located within the virion.
Sections (FFPE) will be enzymatically digested with 0.05% protease XIV in
TBS according to our standard protocol. Would a permeabilization agent
increase access to the interior proteins? Any suggestions/comments will be
greatly appreciated.


Thank you,

Carla Conway
Western Fisheries Research Center
6505 NE 65th St
Seattle, WA 98115
ph: 206-526-6282 ext. 242
fax: 206-526-6654





From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 10:59:08 2005



From: sghoshro-at-NMSU.Edu
Date: Thu, 3 Mar 2005 10:05:24 -0700 (MST)
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The timing of Bill's e-mail to the list serve couldn't be better. Last
week my university (a relatively small institution) administrators brought
a proposal on the table that the EM Lab is too expensive and costing too
much to the university. Being in nowhereland, we hardly get outside users
and we have a fairly small user base. However, it is the one and only EM
facility on campus that serves both materials and biological users. The
administrators did agree that this facility helps them to sell the
university for prospective students and faculty. However, as many people
mentioned that shortsightedness in their part is playing a big role here.

So far we have generated a huge support against this move and I hope it
will help to maintain the lab. We generate a small amount of revenues and
we have one TEM and one SEM(with EDS) and an upright fluorescence scope.
The lab has two staff (one full time, one part time) members and their
salaries come from the college Arts & Sciences and Biology department. The
revenues are mostly used for buying all lab supplies, computer upgrades,
small instrument repairs and occasionally to pay for a student assistant.

We offer a grad level EM course every fall thrpough Biology and is quite
popular among grad students. The course costs $1000.00 every year for
Biology and eight students (max) can enroll due to limited resources.
Department chair mentioned to me that this course is too expensive since
only eight students can take it.

Therefore, my question is to the people who are running small university
facilities if they are facing similar problems from their administrators.

Thanks a lot,

Soumitra


*************************************************************
Soumitra Ghoshroy
College Associate Professor, Biology
Director, Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
http://biology-web.nmsu.edu/ghoshroy/ghoshroy.htm
http://emldata.nmsu.edu/eml/

On Tue, 1 Mar 2005, William P. Sharp wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Microscopy Listers -
}
} Our perennial lament - does anyone out there know of a University or
} Institute (non-profit) that operates an EM facility or a Bioimaging facility
} that takes in enough money through charge back of users to cover most or all
} of the costs of lab operation, to include service contracts on major
} instruments? If so, may we know the rates charged? Are there a relatively
} large number of outside users that are billed at a much higher rate than
} inside users?
}
} We at ASU are in the second year of ramping up the rates charged for zero to
} an average charge rate over four years and are feeling some pressure to
} charge enough to pay for all operations of our bioimaging facilities, EM and
} light as soon as possible. Those of us who are responsible for the facilities
} are concerned about pricing ourselves out of the market and we need some idea
} of whether it can be done without a total upheaval of service or the loss of
} our core facilities.
}
} Thanks for any information or advice you may have for us -
}
} Bill Sharp
}
} William P. Sharp
} Arizona State University
} School of Life Sciences, box 4501
} Tempe, AZ 85287-4501
} Phone - (480)-965-3210
} Fax - (480)-965-6899
}
}


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 11:00:19 2005



From: Hong Yi :      hyi-at-emory.edu
Date: Thu, 3 Mar 2005 12:07:24 -0500
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Everyone:

My facility has been doing relatively well in being self-sufficient
for the last 5 years. So I was surprised when a friend of mine in
Florida told me that her facility is 100% self-sustained. Even more
stunning to me was that her rate in general was lower than mine. She
had to recover everything including salaries, supplies, service
contracts etc. She told me her trick was to charge for everything.
She even charged for time she spent to email users. She also told me
that she does not have any complain from users about her rate.

I joked to her that she is someone I do not want my dean to know
about. ;-).

HOng
Emory EM



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 11:02:02 2005



From: Alan Nicholls :      nicholls-at-uic.edu
Date: Thu, 03 Mar 2005 11:20:03 -0600
Subject: [Microscopy] MMMS Meeting March 24th, 2005

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} From a fair competition point of view, I am glad to hear that most if
not all university facilities increase their fees to outside users. And
70% to 100% increases in your standard rates sounds like a lot. But
when the standard in house rates are on the order of $40 to $50 dollars
per hour for TEM or SEM/EDS (those are university in house rates I have
heard about in the Boston area) then the outside user rates still end up
being about half of what the norm for private microscopy labs is in my
area. And that could tempt even the most loyal of our clients.

But as someone else noted, all is not lost for those of us with the
private labs since we can compete in other areas (service, turn around
times, quality, advertising, location, etc.). And frankly right now
there is more work out there, if you know where to look, then the
university labs can handle, much more. And I do not think doing outside
work is what the staff at university labs want to do all day anyway. So
the overall impact to private sector labs, like mine, is likely much
less we often think.

Sorry if I sounded hostile in my last post. It is just that you guys
get so many nice toys to play with, and for free. But then again I do
not have to deal with any administration types or write grant proposals.
Guess the grass is always greener...

John Knowles
MicroVision Laboratories, Inc.



-----Original Message-----
} From: Leona Cohen-Gould [mailto:lcgould-at-med.cornell.edu]
Sent: Thursday, March 03, 2005 9:27 AM
To: John Knowles; 'Debby Sherman'; 'William P. Sharp';
Microscopy-at-microscopy.com

Local MSA, MAS affiliate meeting announcement

Imaging Biomolecular Interactions
Midwest Microscopy and Microanalysis Society
Co-sponsor: Biological Imaging Facility, Northwestern University

Thursday, March 24, 2005

Northwestern University
Pancoe Life Sciences Building
Pancoe/ENH Auditorium
Evanston, IL 60208

Meeting Schedule
Morning Session
9:00 Registration
9:45 Welcome and introduction
10:00 Victoria Froelich, University of Texas Health Sciences Center San
Antonio: FRET, dectection of molecular interactions.
11:00 Kanya Rajangam, Northwestern University: Cells in self-assembling gels.
11:30 Carina Holmberg and Heather Brignull, Northwestern University:
Imaging-based characterization of aggregates in neurodegenerative disease
models.
12:00-12:45 Lunch
12:45 Business Meeting

Afternoon session
1:00 H. Peter Lu, Paci. c Northwest National Laboratory: Single-molecule
biophysics.
2:00 Debbie Klos, Northwestern University: Recognition of membrane protein
cargo by the Rsp5 ubiquitin ligase.
2:30 Victoria Wu, Northwestern University: The expansive tube, the role of
the septate junction in Drosophila tracheal development.
3:00 Reception and Student Poster Competition Exhibit and Judging.

RSVP Required
Please send RSVP via email, phone, or fax to:
Arvid Casler, MMMS Program Coordinator
c/o MAI
P. O. Box 394
Mundelein, IL 60060
phone/FAX: (847) 566-7716
email: arvid_casler-at-fmo.com


Admission: Free to MMMS Members
MMMS Membership: $10.00
MMMS membership available at registration

Driving directions to Northwestern University Evanston Campus available on
the Web.
From the north or northwest, via Interstate 88, Interstate 90 or
Interstate 190
(Note: these are the directions to follow if traveling from O'Hare
International Airport)
http://www.northwestern.edu/campus/directions/evanston-north-northwest.html
From the west, via Interstate 94
http://www.northwestern.edu/campus/directions/evanston-west.html
From the west or southwest, via Interstate 55 or Interstate 80
http://www.northwestern.edu/campus/directions/evanston-west-southwest.html
From the south or southeast, via Interstate 94, Interstate 90, Interstate
80 or Interstate 57
http://www.northwestern.edu/campus/directions/evanston-south-southeast.html

PARKING
Parking permits are $4 and are nonrefundable.
Parking can be challenging on the NU Evanston Campus especially after 9AM.
Spaces in selected lots are available on a first-come-first-served basis.



Alan W Nicholls, PhD
Director of Research Service Facility (Electron Microscopy)
Research Resources Center - East (M/C 337)
Room 100 Science and Engineering South Building
The University of Illinois at Chicago
845 West Taylor St
Chicago, IL 60607-7058

Tel: 312 996 1227
Fax: 312 996 8091
Office: Room 110

Web site www.rrc.uic.edu



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 11:24:13 2005



From: Aleksandr Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Thu, 03 Mar 2005 17:34:01 +0000
Subject: [Microscopy] Re: TEM EM Atlas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thank you everybody for the information about EM Atlas!

It seems that the most popular EM atlases are the ones by Fawcett and by
Johannes A G Rhodin.

I will try to find them.

--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 11:28:23 2005



From: Aleksandr Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Thu, 03 Mar 2005 17:38:12 +0000
Subject: [Microscopy] TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear collegues!

Just recently I found the structure, which I cannot identify. It has
regular hexagonal pattern and localized in the cytosol of the cells in
embryonic tendons. It is not membrabous, some kind of cytoskeletal?

Could you, please, take a look and hint what it could be?
The address for pictures is (structure is identified by arrows):
http://sysoj.spymac.net/image.jpg
http://sysoj.spymac.net/image2.jpg

Regards,
--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 11:47:08 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Thu, 3 Mar 2005 09:53:16 -0800
Subject: [Microscopy] RE: Re: RE: Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


John;
You are right to be dismayed. Companies are so cost obsessed
that they will do anything to save a buck, and that includes getting as
much for nearly free as they can. My own company, a $30 billion Fortune
500 company, pushed for using university labs for years, just to avoid
the cost of purchasing, housing, staffing, and maintaining a TEM,
something they could easily afford. Some people just have no shame.

John Mardinly

} Bill et al,
}
} I own and operate a small analytical microscopy lab and like everyone
} else in the dreaded private sector, I have to make a profit every
month
} or else I do not get paid. Because of this my focus is naturally on
} serving the client's microscopy needs and not so much on research.
}
} I was always under the impression that Universities were mainly
focused
} on research. When part of my tax dollars goes to pay for new
equipment
} for University microscopy labs I am ok with it as long as I believe
} that
} it is going to be used for education and/or research projects that the
} University is doing. When I hear that this equipment is being used to
} set up microscopy service labs that are competing against me then I am
} definitely not ok with it. Why should I have to pay for all of my
} equipment, come in on weekends and stay late into the night doing my
} own
} equipment maintenance and service, and pay 100% of my costs when the
} government is using my tax money to help my competition?
}
} There are a number of local Universities in the Boston area that are
} happy to accept government grants for new equipment and then use it to
} take business away from privately owned labs like mine. I have talked
} to many companies that have told me that they go to a University
} Microscopy lab to get there work done instead of a lab like mine
} because
} the Universities rates are so low. If I only had to cover 80% (or
} less)
} of my costs and got grants for my equipment I could drop my rates as
} well. MIT is one of the few Universities that play by the rules.
They
} do not take business away from privately own labs like mine. In fact
} they send private companies away and recommend labs like mine to them.
} And for that I am grateful.
}
} I do not think Universities should compete against private labs when
} they are getting government money (just in case you have not figured
} that out by now 8^). If they must then at least raise your rates to
} the
} typical rates in your area. That would make it a fair playing field.
}
} Another thought I had, do your administrators factor in any of the
} students tuition when calculating your profitability? There are many
} students that pick which college they are going to attend based on, at
} least in part, what the facilities are at that location. A microscopy
} department is a big asset to a school that is trying to attract
} students
} interested in the sciences (e.g. biology and materials). When you
look
} at a University Microscopy Dept. as a part of the whole education
} system
} then it is silly to think it has to act like a business. The campus
} grounds need to be kept up as well in order for a university to
attract
} and keep students. Does your administration ask the grounds crew to
be
} profitable as well? Maybe compete with locate landscaping companies
} mowing lawns and trimming bushes. I bet not.
}
} John Knowles
} President
} MicroVision Laboratories, Inc.
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446





From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 13:02:38 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Thu, 03 Mar 2005 14:09:02 -0500
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} She sounds like a lawyer....



} Dear Everyone:
}
} My facility has been doing relatively well in being
} self-sufficient for the last 5 years. So I was surprised when a
} friend of mine in Florida told me that her facility is 100%
} self-sustained. Even more stunning to me was that her rate in
} general was lower than mine. She had to recover everything
} including salaries, supplies, service contracts etc. She told me
} her trick was to charge for everything. She even charged for time
} she spent to email users. She also told me that she does not have
} any complain from users about her rate.
}
} I joked to her that she is someone I do not want my dean to
} know about. ;-).
}
} HOng
} Emory EM


--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 13:06:26 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Thu, 03 Mar 2005 14:12:52 -0500
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Aleksandr...
your dissection was a little mis-directed. Those are thin and thick
muscle filaments in cross section. Your sample must have come from
the point of muscle-tendon junction.
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 14:07:57 2005



From: Jay Campbell :      microtomy-at-gmail.com
Date: Thu, 3 Mar 2005 14:14:11 -0600
Subject: [Microscopy] LASIK correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello microscopists,
A quick check of the archives indicates that it has been almost 4
years since there was much discussion about this topic on list. Given
the pace of technology and the popularity of the proceedure, it seems
fitting to visit the topic again. I have recently been evaluated and
declared an excellent candidate for LASIK correction. I was hoping
that others who've opted for this surgery could share their
experience, positive or negative, to help me make my decision.

Thanks much and I will be happy to post a compilation of the responses,
Jay


Jay Campbell
Research Specialist
University of Wisconsin
Laboratory of Molecular Biology
R.M. Bock Labs
1525 Linden Drive
Madison, WI 53706
jmcampbe-at-wisc.edu
608 263 8481


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 14:48:14 2005



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Thu, 03 Mar 2005 15:54:46 -0500
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I'll take a stab at it and say it looks like virus in a crystalline array
although size appears a bit small.

Is there any signs of fibers in longitudinal orientation? If it is made up
of cytoskeletal elements, i.e. fibers of different diameters, than I would
expect to see some in longitudinal as well as x-sectional view.

Debby

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy


On 3/3/05 12:38 PM, "Aleksandr Mironov" {Aleksandr.Mironov-at-manchester.ac.uk}
wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Dear collegues!
}
} Just recently I found the structure, which I cannot identify. It has
} regular hexagonal pattern and localized in the cytosol of the cells in
} embryonic tendons. It is not membrabous, some kind of cytoskeletal?
}
} Could you, please, take a look and hint what it could be?
} The address for pictures is (structure is identified by arrows):
} http://sysoj.spymac.net/image.jpg
} http://sysoj.spymac.net/image2.jpg
}
} Regards,




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 15:06:28 2005



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Thu, 03 Mar 2005 15:19:14 -0600
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

aleksandr

it is not the best way to look at these things, all i have is a picture,
no size bars, etc. also, are the cells tissue culture, are they an
explant, or are they harvested tissue which you have embedded?

having said those things, looking at the paracrystaline nature makes me
think virus. have you considered lysing the cells and looking at the
gemisch by negative stain? that would confirm or reject the possibility
of virus.

paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-954
Cell:204-781-1502
Fax:204-789-3926



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 15:58:28 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Thu, 03 Mar 2005 14:08:06 -0800
Subject: [Microscopy] Re: RE: Re: Cost Recovery in Imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

John
Very good point. I think, the University's facilities (including EM)
should serve scientific community and stay away from the business (I meant
business of "generating" money). In my prospective, "business" orientation
of the University Labs seriously affected the quality of research: they
tends to use "standard" simple protocols and users don't want to pay for
extra hour to develop better protocol for their particular case. I think,
most of us here at ListServer do understand this perfectly. The problem is
that Universities are struggled from money deficit and we are a good
targets to blame: we are using too much money... not "profitable"...
etc; how you could measure "profit" in science?). To be truthful, I have
to admit that I don't feel such pressure at UCLA and I met full
understanding that I am serving Science. Therefore my facility is heavily
subsidized. I also have rates for outsiders 6x higher than for insiders -
mostly to keep them away, so I have more time for students etc. Rates for
insiders are diversified, so many users use instruments for free. I clearly
understand that my "good life" may ended any moment and I'll join "the
club" switching from doing science to searching for the money... Have a
great day, Sergey

At 05:49 PM 3/2/2005 -0500, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 16:04:47 2005



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Thu, 03 Mar 2005 16:17:33 -0600
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

aleksandr

another possibility is that you have actin/myosin filaments in muscle
fibres in cross section. it has been a long time since i've looked at
smooth muscle but i don't think it would match for that. it would not
fit for the standard cross section of a skeletal muscle cell, i would
expect more ordered fibres. but i do not know what the fibres would
look like in a myocyte at the edge of an embryonic tendon. again,
problem is lack of size markers makes it difficult

paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-954
Cell:204-781-1502
Fax:204-789-3926



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 16:39:53 2005



From: stuart McKernan :      stuartm-at-umn.edu
Date: Thu, 3 Mar 2005 16:46:52 -0600
Subject: [Microscopy] LAS meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The Minnesota Microscopy Society will be co-hosting the 3rd Annual
Minnesota Joint Technical Symposium
(MinnTS 2005) on Thursday, March 17, 2005. Any interested
microscopists will be welcome to attend. This year the topic is
Biomaterials and Biotechnology, and we will have two speakers:

Vincent Chow , Optobionics, www.optobionics.com
"Artificial Retina"
and
Arthur J. Coury, Genzyme Corporation,
"Therapeutic Strategies, from Replacement Parts to Regenerative
Medicine: Challenges and Opportunities"

The Schedule for the afternoon is as follows
Schedule of Events
4:30 - 5:30 Registration, tours, and social
5:30 - 6:40 Dinner
6:45 - 7:00 Welcome message
7:00 - 7:50 Vincent Chow
7:50 - 8:00 Break
8:05 - 8:55 Arthur Coury

The cost for the event will be $25 including dinner. Reservations
should be addressed to: Bede Willenbring, (651) 236-5430 or
Bede.Willenbring-at-HBFuller.com by March 11, 2005.

Please include your name and company, email address and phone number
with your registration, as well as whether you are a member of any of
the sponsoring societies:
American Chemical Society, Minnesota chapter (ACS)
American Institute of Chemical Engineers, Upper Midwest Section
(AIChE)
American Society for Materials, Minnesota chapter (ASM)
American Vacuum Society, Minnesota Chapter (AVS)
Minnesota Microscopy Society (MMS)
Society for Applied Spectroscopy, Minnesota chapter (SAS)
SEMI-MN
Society for Information Display, Midwest chapter (SID)

More details can be found on our website http://www.MNmicroscopy.org


Prof. Stuart McKernan
IT Characterization Facility, University of Minnesota            
E-mail :  stuartm-at-umn.edu
12 Shepherd Labs,                                                    
                 Office:         (612) 624-6009
100 Union Street S. E., Minneapolis, MN 55455                    
Lab:            (612) 626-7594





From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 17:38:44 2005



From: Schmitz, Bob :      rschmitz-at-uwsp.edu
Date: Thu, 3 Mar 2005 17:45:41 -0600
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Since you indicate that this is embryonic tendon, is there any chance that these cells are not fibroblast like but muscle like? Those hexagonal arrays look similar to the cross-sections of skeletal muscle myofibrils where the thick and thin filaments overlap.

Bob
Robert J. Schmitz
Department of Biology
University of Wisconsin Stevens Point
Stevens Point, WI 54481
715-346-2420
Email: rschmitz-at-uwsp.edu
http://biology.uwsp.edu/faculty/RSchmitz/Default.htm



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 18:11:27 2005



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Thu, 03 Mar 2005 19:17:59 -0500
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Surely these must be myofibrils, in cross section.
A cell differentiating into a myocyte maybe,
which could be expected in embryonic tissue.
Or a smooth muscle cell?
I would opt for the former personally.

Best wishes

Marc

On Thursday, March 3, 2005, at 12:38 PM, Aleksandr Mironov wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Dear collegues!
}
} Just recently I found the structure, which I cannot identify. It has
} regular hexagonal pattern and localized in the cytosol of the cells in
} embryonic tendons. It is not membrabous, some kind of cytoskeletal?
}
} Could you, please, take a look and hint what it could be?
} The address for pictures is (structure is identified by arrows):
} http://sysoj.spymac.net/image.jpg
} http://sysoj.spymac.net/image2.jpg
}
} Regards,
} --
} Aleksandr Mironov
} Experimental Officer
} G450A, Stopford Building
} EM Unit, Faculty of Life Sciences
} University of Manchester
} Oxford Road
} Manchester
} M13 9PT
} UK
}
} Tel. +44-(0)161-275-7395
} Fax. +44-(0)161-275-5171
} E-mail: Aleksandr.Mironov-at-manchester.ac.uk
} MSN: amironov-at-hotmail.co.uk
} Web: http://www.empgu.man.ac.uk
}
}
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 3 21:50:20 2005



From: Ken Tiekotter :      tiekotte-at-up.edu
Date: Thu, 03 Mar 2005 19:56:54 -0800
Subject: [Microscopy] RE: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

HAMLET Do you see yonder cloud that's almost in shape of a camel?
POLONIUS By th'mass, and 'tis: like camel, indeed.
HAMLET Methinks it is like a weasel.
POLONIUS It is backed like a weasel.
HAMLET Or like a whale.
POLONIUS Very like a whale.
William Shakespeare, Hamlet, c. 1600 [Act 3, scene 2]

I must chuckle. Have you looked at other cells and found these structures
in different planes of view, hence allowing better differentiation?

The n of 1 makes for a tough puzzle.

:-) Ken

_______________________________________
Kenneth L. Tiekotter, Adjunct Professor
The University of Portland
Department of Biology
5000 N Willamette Blvd.
Portland, OR 97203 USA

Tel.: 503.493.8861






From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 01:39:26 2005



From: Gareth Morgan :      Gareth.Morgan-at-labmed.ki.se
Date: Fri, 04 Mar 2005 08:46:46 +0100
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

Look like rhabdomyocyte/blast actin/myosin contractile fibres to me - would
be seen in that typical array. I don't think virus (but then I am not a
virologist :-)).

A scale would be nice but I can see the problem really - there are enough
identifiable structures to be able to 'size' the array. Might be different
if I was going to court about it!

Good luck with your hunt for Bloom and for Rhodin - I still have my old
copies - wonderful books.


Gareth


With best wishes,

Gareth

http://www.ki.se/biomedlab
e-mail Gareth.Morgan-at-labmed.ki.se

Tel +46 8 5858 1038
Fax +46 8 5858 7730

Gareth Morgan MPhil MSc FIBMS,
Department of Histo/cytopathology, Laboratory Medicine (Labmed),
Karolinska Institute,
Karolinska University Hospital at Huddinge, F46
SE 141 86 Stockholm
Sweden

OBS! Besöksadress: F-Huset, Forskningsgatan 2 F52, Rum 2.10. Laboratoriet
för klinisk patologi och cytologi.

NB! Visiting address: Building F, Research Corridor 2 F52, Room 2.10.
Clinical Histo- and Cytopathology Laboratory.




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 02:06:51 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Fri, 4 Mar 2005 00:12:40 -0800
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Soumitra;
Remind your administrators about the super conducting
supercollider. It was cancelled, and now virtually all of the high
energy physics research (and all the physicists) is done in Europe!

John Mardinly
Intel

-----Original Message-----
} From: sghoshro-at-NMSU.Edu [mailto:sghoshro-at-NMSU.Edu]
Sent: Thursday, March 03, 2005 9:05 AM
To: William P. Sharp
Cc: Microscopy-at-microscopy.com

The timing of Bill's e-mail to the list serve couldn't be better. Last
week my university (a relatively small institution) administrators
brought
a proposal on the table that the EM Lab is too expensive and costing too

much to the university. Being in nowhereland, we hardly get outside
users
and we have a fairly small user base. However, it is the one and only EM

facility on campus that serves both materials and biological users. The
administrators did agree that this facility helps them to sell the
university for prospective students and faculty. However, as many people

mentioned that shortsightedness in their part is playing a big role
here.

So far we have generated a huge support against this move and I hope it
will help to maintain the lab. We generate a small amount of revenues
and
we have one TEM and one SEM(with EDS) and an upright fluorescence scope.

The lab has two staff (one full time, one part time) members and their
salaries come from the college Arts & Sciences and Biology department.
The
revenues are mostly used for buying all lab supplies, computer upgrades,

small instrument repairs and occasionally to pay for a student
assistant.

We offer a grad level EM course every fall thrpough Biology and is quite

popular among grad students. The course costs $1000.00 every year for
Biology and eight students (max) can enroll due to limited resources.
Department chair mentioned to me that this course is too expensive since

only eight students can take it.

Therefore, my question is to the people who are running small university

facilities if they are facing similar problems from their
administrators.

Thanks a lot,

Soumitra


*************************************************************
Soumitra Ghoshroy
College Associate Professor, Biology
Director, Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Tel: 505-646-3268 (office), 646-3283 (lab)
Fax: 505-646-3282
e-mail:sghoshro-at-nmsu.edu
http://biology-web.nmsu.edu/ghoshroy/ghoshroy.htm
http://emldata.nmsu.edu/eml/

On Tue, 1 Mar 2005, William P. Sharp wrote:

}
}
}
------------------------------------------------------------------------
------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe --
http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
-------
}
} Microscopy Listers -
}
} Our perennial lament - does anyone out there know of a University or
} Institute (non-profit) that operates an EM facility or a Bioimaging
facility
} that takes in enough money through charge back of users to cover most
or all
} of the costs of lab operation, to include service contracts on major
} instruments? If so, may we know the rates charged? Are there a
relatively
} large number of outside users that are billed at a much higher rate
than
} inside users?
}
} We at ASU are in the second year of ramping up the rates charged for
zero to
} an average charge rate over four years and are feeling some pressure
to
} charge enough to pay for all operations of our bioimaging facilities,
EM and
} light as soon as possible. Those of us who are responsible for the
facilities
} are concerned about pricing ourselves out of the market and we need
some idea
} of whether it can be done without a total upheaval of service or the
loss of
} our core facilities.
}
} Thanks for any information or advice you may have for us -
}
} Bill Sharp
}
} William P. Sharp
} Arizona State University
} School of Life Sciences, box 4501
} Tempe, AZ 85287-4501
} Phone - (480)-965-3210
} Fax - (480)-965-6899
}
}




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 03:16:10 2005



From: Gareth Morgan :      Gareth.Morgan-at-labmed.ki.se
Date: Fri, 04 Mar 2005 10:23:45 +0100
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

'can' in my previous mail should of course be 'can't' - corrected version below


Hi

Look like rhabdomyocyte/blast contractile fibres to me - would be seen in
that typical array. I don't think virus (but then I am not a virologist :-)).

A scale would be nice but I can't see the problem really - there are enough
identifiable structures to be able to 'size' the array. Might be different
if I was going to court about it.

Good luck with your hunt for Bloom and for Rhodin - I still have my old
copies - wonderful books.

Gareth

With best wishes,

Gareth

http://www.ki.se/biomedlab
e-mail Gareth.Morgan-at-labmed.ki.se

Tel +46 8 5858 1038
Fax +46 8 5858 7730

Gareth Morgan MPhil MSc FIBMS,
Department of Histo/cytopathology, Laboratory Medicine (Labmed),
Karolinska Institute,
Karolinska University Hospital at Huddinge, F46
SE 141 86 Stockholm
Sweden

OBS! Besöksadress: F-Huset, Forskningsgatan 2 F52, Rum 2.10. Laboratoriet
för klinisk patologi och cytologi.

NB! Visiting address: Building F, Research Corridor 2 F52, Room 2.10.
Clinical Histo- and Cytopathology Laboratory.




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 04:13:48 2005



From: Aleksandr Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Fri, 04 Mar 2005 10:23:15 +0000
Subject: [Microscopy] Re: Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thank you a lot for your advices!

Sorry, that I do not have more decent image to show and that I forgot to
put a scale bar.

Two suggestions were made:
1) It is a virus
2) It is a forming myofibrilles in embryonic myocyte

We have checked serial sections from the sample (tail from mouse embryo)
and it seems that this structure goes through several of them. The cell
containing the strucutre is located just nearby the tendon. So, the most
probable variant is that these are myofibrilles.

Thank you very much again!
Certainly I need to find some good EM Atlas to educate myself better :-)

--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 04:37:37 2005



From: shashi singh :      shashis_99-at-yahoo.com
Date: Fri, 4 Mar 2005 02:44:16 -0800 (PST)
Subject: [Microscopy] internal structure by TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I feel they may be ribosomes. This would be a period
of high protein synthesis and ribosomes can form such
arrays in many tissues.
ANY prizes for correct guessing!

Shashi Singh
CCMB
Hyderabad
INDIA

=====
Shashi Singh
Scientist
Centre for Cellular and Molecular Biology
Hyderabad-500 007
INDIA
PH-91-40-7192575,7192761,7192615
FAX-91-40-7160591, 7160311




__________________________________
Celebrate Yahoo!'s 10th Birthday!
Yahoo! Netrospective: 100 Moments of the Web
http://birthday.yahoo.com/netrospective/


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 10:06:15 2005



From: Franklin Bailey :      jfb-at-uidaho.edu
Date: Fri, 04 Mar 2005 08:20:27 -0800
Subject: [Microscopy] LASIK correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Jay,
I had the correction operation in 2001 and would do it again in a heartbeat.
Now the only correction I need is the typical old folks reading glasses.
Since I was wearing bifocals before the procedure, reading glasses were a
given. It's great to be able to look outside and see the horizon in clear
and sharp focus.

Franklin Bailey
Electron Microscopy Center
University of Idaho
Moscow, ID 83844-2204

-----Original Message-----
} From: Jay Campbell [mailto:microtomy-at-gmail.com]
Sent: Thursday, March 03, 2005 12:14 PM
To: Microscopy-at-microscopy.com

Hello microscopists,
A quick check of the archives indicates that it has been almost 4
years since there was much discussion about this topic on list. Given
the pace of technology and the popularity of the proceedure, it seems
fitting to visit the topic again. I have recently been evaluated and
declared an excellent candidate for LASIK correction. I was hoping
that others who've opted for this surgery could share their
experience, positive or negative, to help me make my decision.

Thanks much and I will be happy to post a compilation of the responses,
Jay


Jay Campbell
Research Specialist
University of Wisconsin
Laboratory of Molecular Biology
R.M. Bock Labs
1525 Linden Drive
Madison, WI 53706
jmcampbe-at-wisc.edu
608 263 8481



From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 10:15:42 2005



From: Franklin Bailey :      jfb-at-uidaho.edu
Date: Fri, 04 Mar 2005 08:30:03 -0800
Subject: [Microscopy] TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Many years ago when I was at the University of Texas Medical Branch in
Galveston working for Dr. Ward Kischer, we did extensive research on scar
tissue and the granulation tissue of burned skin. Based on the experience
of looking at thousands of sections of connective tissue, I believe the
cells to actually be fibroblasts and the particles to be cross sections of
collagen fibrils, which can range anywhere from 10 to over 200 nm in
diameter, and are produced by the fibroblasts.

Franklin Bailey
Electron Microscopy Center
University of Idaho
Moscow, ID 83844-2204

-----Original Message-----
} From: Aleksandr Mironov [mailto:Aleksandr.Mironov-at-manchester.ac.uk]
Sent: Thursday, March 03, 2005 9:38 AM
To: Microscopy-at-microscopy.com

Dear collegues!

Just recently I found the structure, which I cannot identify. It has
regular hexagonal pattern and localized in the cytosol of the cells in
embryonic tendons. It is not membrabous, some kind of cytoskeletal?

Could you, please, take a look and hint what it could be?
The address for pictures is (structure is identified by arrows):
http://sysoj.spymac.net/image.jpg
http://sysoj.spymac.net/image2.jpg

Regards,
--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk





From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 10:19:01 2005



From: Evren Cubukcu :      ecubukcu-at-hacettepe.edu.tr
Date: Fri, 04 Mar 2005 18:25:54 +0200
Subject: [Microscopy] SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi there;
We have been trying to decide which SEM is worth buying.
JEOL JSM-6380 LV (with PGT Avalon AV 8000 - Sahara EDS system)
or
CARL ZEISS (LEO) EVO 40 XVP (with Quantax, QX2 EDS system)
The instrument will be used for geological purposes and most of the
specimens will be rocks.
We think that EVO 40 XVP is better with motorized 5 axis, no water
cooling, turbomolecular pump... And Quantax EDS (Rontec x-flash
dedector) is amazing being standartless and LN free and over 400,000
cps.
I really need to hear comments on both instruments from users.

Thank You

Evren Cubukcu
Hacettepe University
Dept. Geol. Eng.
SEMProbe Lab.
Ankara
Turkey




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 11:25:07 2005



From: michael shaffer :      michael-at-shaffer.net
Date: Fri, 4 Mar 2005 14:02:05 -0330
Subject: [Microscopy] automatic dessicating cabinets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Is anyone familiar with the automatic dessicating cabinets available ...
e.g., "SECADOR" cabinets which are available from Electron Microscopy
Sciences. How do they work? More importantly, how well do they work?

tia & cheerios ... shAf :o)
Avalon Peninsula, Newfoundland
{www.micro-investigations.com}



From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 11:50:15 2005



From: White, Woody N. :      NWWhite-at-bwxt.com
Date: Fri, 4 Mar 2005 12:55:39 -0500
Subject: [Microscopy] SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I don't have experience with either system so have only one comment.
For geological work, I would expect light element capability would be
important to you. If so, be sure to carefully evaluate the low energy
sensitivity and resolution of LN2 free detectors.

Woody



-----Original Message-----
} From: ecubukcu-at-hacettepe.edu.tr [mailto:ecubukcu-at-hacettepe.edu.tr]
Sent: Friday, March 04, 2005 11:26 AM
To: Microscopy-at-microscopy.com

Hi there;
We have been trying to decide which SEM is worth buying.
JEOL JSM-6380 LV (with PGT Avalon AV 8000 - Sahara EDS system)
or
CARL ZEISS (LEO) EVO 40 XVP (with Quantax, QX2 EDS system)
The instrument will be used for geological purposes and most of the
specimens will be rocks.
We think that EVO 40 XVP is better with motorized 5 axis, no water
cooling, turbomolecular pump... And Quantax EDS (Rontec x-flash
dedector) is amazing being standartless and LN free and over 400,000
cps.
I really need to hear comments on both instruments from users.

Thank You

Evren Cubukcu
Hacettepe University
Dept. Geol. Eng.
SEMProbe Lab.
Ankara
Turkey






From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 12:46:09 2005



From: john hoffpauir :      hoffpajo-at-yahoo.com
Date: Fri, 4 Mar 2005 10:52:40 -0800 (PST)
Subject: [Microscopy] Re: LASIK correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

i personaly prefer contacts over a laser in the eye.
better than a sharp stick i guess.
--- Jay Campbell {microtomy-at-gmail.com} wrote:
}
}
}
------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The
} Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help
}
http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
-------------------------------------------------------------------------------
}
} Hello microscopists,
} A quick check of the archives indicates that it has
} been almost 4
} years since there was much discussion about this
} topic on list. Given
} the pace of technology and the popularity of the
} proceedure, it seems
} fitting to visit the topic again. I have recently
} been evaluated and
} declared an excellent candidate for LASIK
} correction. I was hoping
} that others who've opted for this surgery could
} share their
} experience, positive or negative, to help me make my
} decision.
}
} Thanks much and I will be happy to post a
} compilation of the responses,
} Jay
}
}
} Jay Campbell
} Research Specialist
} University of Wisconsin
} Laboratory of Molecular Biology
} R.M. Bock Labs
} 1525 Linden Drive
} Madison, WI 53706
} jmcampbe-at-wisc.edu
} 608 263 8481
}
}




__________________________________
Celebrate Yahoo!'s 10th Birthday!
Yahoo! Netrospective: 100 Moments of the Web
http://birthday.yahoo.com/netrospective/


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 13:42:08 2005



From: Daniel Geiger :      geiger-at-vetigastropoda.com
Date: Fri, 4 Mar 2005 11:48:32 -0800
Subject: [Microscopy] Re: SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Evren,

we just installed in January an Zeiss EVO 40 XVP and went through he
same evaluation process. We use it mainly for biological samples
many of them uncoated, but no EDS or similar application.

The main advantages I see for the EVO are:
- No need for water cooling
- True continuous zoom control of magnification (not stepped: has
driven me crazy on other instruments).
- Most other adjustments are also truly continuous (e.g., EHT, spot
size/probe current, detector bias, VP), not stepped.
- large image files (3200 x 2300 pixels, 6.7 MP)
- Signal mixing (e.g., continuous mixing of SE/VPSE signal with
selected quadrants of QBSD detector): that is REALLY helpful
particularly with uncoated specimen.

Low mag is a bit limited for VP mode/fixed aperture in place with
about 10-15 mm specimen being the largest to be imaged, depending on
WD and ETH. In HiVac, though you can get up to about 25 mm.

I have used a Hitachi 3000N with ESED detector, and the Zeiss VPSE
detector is by far better. If I recall correctly, JEOL does not
produces a variable pressure/environmental SE detector. The number of
options for noise reduction and dealing with charging are very nice.

Re motorized stage, the five axes are great, though the tilt lever is
set-up backwards: the + tilt is pushing to the right, the computer
screen icon tilts to the right, but the stage actually tilts to the
left. So when you visualize the stage tilt using the IR scope (which
is at the front of the chamber, so you get a non-inverted image of
the chamber) you get a bit cross-eyed. I trust Zeiss can fix that on
your instrument.

Get the extended desktop option, so you can use one mouse to go to
two screens. There is also an option for having two live images (say
SE and BSD). The last cool thing is the "$7000 keyboard". It is a
custom keyboard with the main controls as knobs (mag, focus,
stigmation, gun tilt, gun shift, image rotate, scan speed, reduce
raster, brightness contrast) The actual keyboard is like that of a
laptop computer, i.e. without numerical keypad. It is a great time
saver and you may avoid repetitive stress problems when using a mouse
for a long time. I've done some student training, and I think the
keyboard is a bit more intuitive than screen controls so it speeds
up the learning process.

Last but not least, the simplicity of the instrument is awe
inspiring. There are just 5 computer boards in the machine, and very
few cables running around the back. I think my desktop computer with
printer and network connection has more cables than the SEM!


Obviously, I am a happy user. Just for the record, I have no
interests in the Carl Zeiss company other then being a heavy user of
their products (photo, LM, SEM).

Best wishes Daniel


At 6:25 PM +0200 3/4/05, Evren Cubukcu wrote:
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
***************************************************************************************

Daniel L. Geiger, Ph.D.
Invertebrate Zoology
Santa Barbara Museum of Natural History
2559 Puesta del Sol Road
Santa Barbara, CA 93105

voice (805) 682 4711 x152

e-mail geiger-at-vetigastropoda.com
www http://www.vetigastropoda.com


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 14:06:13 2005



From: Anjeanette Ormonde :      Anjeanette.Ormonde-at-unilever.com
Date: Fri, 4 Mar 2005 15:12:52 -0500
Subject: [Microscopy] Metal Corrosion Reference

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Can anybody recommend a good reference that will help me figure out what
different types of corrosion look like? In the past I've been given samples of
steel containers where corrosion was suspected and was able to identify it
using EDS. This time there is a layer of steel coated with a thin layer of
tin. After storage with a solution in it a portion of the can is slightly
discolored. I was asked if I could determine if the discoloration is due to
detinning. I can't polish the samples flat but I did get a strong signal for
Tin using EDS so I'm convinced the layer is still there. But, I realize that I
do not know what to look for to determine if it is some sort of tin corrosion.
Any advice you can give on this problem would be greatly appreciated.
In addition, I would like to get a good reference so the next time I'm
presented with a new type of corrosion I can use it to help me figure out what
is going on.

Thanks,
Angie


Anjeanette Ormonde
Sr. Project Scientist - Unilever HPC
Rolling Meadows, IL 60008
Ph: 847-734-3539


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 4 14:22:29 2005



From: gay.ribisi-at-ns.microscopy.com (by way of MicroscopyListserver)
Date: Fri, 4 Mar 2005 18:09:14 -0600
Subject: [Microscopy] viaWWW: microscope with camera attachment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Jay,

I also had the surgery in 2001 and would do it all again too. Money well
spent!

Al Coritz
Electron Microscopy Sciences
www.emsdiasum.com



----- Original Message -----
} From: "Franklin Bailey" {jfb-at-uidaho.edu}
To: "Jay Campbell" {microtomy-at-gmail.com} ; {Microscopy-at-microscopy.com}
Sent: Friday, March 04, 2005 11:20 AM

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (gay ribisi) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday, March 4, 2005 at 13:42:01
---------------------------------------------------------------------------

Email: gay ribisi
Name: Gay Ribisi

Organization: artist

Title-Subject: [Microscopy] [Filtered] MListserver: microscope with camera attachment

Question:
I am trying to find a low cost dissecting stereomicroscope with a camera port to attach my Nikon f100 or Nikon D100. Any suggestions?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sat Mar 5 09:51:22 2005



From: Kestutis Smalinskas :      smalinskas-at-yahoo.com
Date: Sat, 5 Mar 2005 07:58:29 -0800 (PST)
Subject: [Microscopy] Re: Metal Corrosion Reference

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I can't think of a reference that would help you in
this case, but I do have a few thoughts I'd like to
pass on.

Firstly, by using EDS methods you can determine if
there is a general loss of Tin on the surface by
looking at the Iron peak. Using a control sample
(either a new container, or better yet, an intact
surface from the same container), and assuming the tin
layer is thin enough that you can pick up iron from
underneath the tin (you may need to go to max kV to
accomplish this), you may see a significant difference
in Iron peak height which would correspond to a
difference in tin layer thickness. Naturally, you'll
want to use the smallest magnification to sample the
largest area for good statistical sampling.

Don't you see any other peaks in the stained area,
when compared with an unstained area (such as Oxygen
or Sulfur)? These other peaks should give you an idea
of what is happening to the stained area.

Secondly, if you have access to an X-ray
diffractometer, you may be able to use powder
diffraction techniques to identify the compounds on
the stained surface. Again, you should run a control
sample (unstained area) for comparison.

Lastly, the stained area can be characterized in a
cross section perpendicular to the surface by using
metallographic methods. To examine thin layers on the
surface, it is common to first electrolytically coat
the sample with nickel to avoid the edge rounding
effect that is common in metallographic samples.
Polishing by conventional metallographic methods would
then give a relatively true and flat view of the tin
layer. The methods would need to be somwhat modified
to prevent corrosion or staining of the cross section
- methods such as avoiding water after 600-grit
polish. Struers has a nice writeup on looking at zinc
layers on steel which would apply in your situation.

Regards,

Stu Smalinskas, P.E.
Sr. Metallurgist
SKF
Plymouth, Michigan
734-414-6862

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

Angie wrote:

Can anybody recommend a good reference that will help
me figure out what different types of corrosion look
like? In the past I've been given samples of steel
containers where corrosion was suspected and was able
to identify it using EDS. This time there is a layer
of steel coated with a thin layer of tin. After
storage with a solution in it a portion of the can is
slightly discolored. I was asked if I could determine
if the discoloration is due to detinning. I can't
polish the samples flat but I did get a strong
signal for Tin using EDS so I'm convinced the layer is
still there. But, I realize that I do not know what
to look for to determine if it is some sort of tin
corrosion.

Any advice you can give on this problem would be
greatly appreciated.

In addition, I would like to get a good reference so
the next time I'm presented with a new type of
corrosion I can use it to help me figure out what
is going on.

Thanks,
Angie


Anjeanette Ormonde
Sr. Project Scientist - Unilever HPC
Rolling Meadows, IL 60008
Ph: 847-734-3539




__________________________________
Celebrate Yahoo!'s 10th Birthday!
Yahoo! Netrospective: 100 Moments of the Web
http://birthday.yahoo.com/netrospective/


From MicroscopyL-request-at-ns.microscopy.com Sat Mar 5 10:04:25 2005



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sat, 05 Mar 2005 11:11:25 -0500
Subject: [Microscopy] Secador Desiccator Cabinets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Michael Shaffer has written:
===================================================================
Is anyone familiar with the automatic dessicating cabinets available ... e.g
, "SECADOR" cabinets which are available from Electron Microscopy Sciences.
How do they work? More importantly, how well do they work?
===================================================================
The Secador™ family of desiccator cabinets are explained in detail (as well
as their operation) on the SPI Supplies website at URL
http://www.2spi.com/catalog/supp/desiccator-cabinets-secador.html

To put it simply, the cabinets have an automatic desiccant regenerating
station. It goes through an automatic cycle where by every so often (about
twenty minutes), the regenerating station is closed off from the volume of
the cabinet, the desiccant is heated, and the moisture is "blown off" to the
outside. After the heat cycle, the doors to the outside are closed, and the
doors to the inside of the chamber are opened, and the process of cabinet
desiccation started all over again.

SPI Supplies has sold these cabinets to a number of customers working not
only in microscopy laboratories but into other settings where one would
prefer to not have to handle particle-generating desiccants.

How well do they work? They work as as well as any other desiccator
cabinet but they work "better" if you factor in the fact that one does not
have to rely on remembering to change desiccant to keep the contents of the
chamber dry.

Disclaimer: SPI Supplies is a distributor for the Secador line of
desiccating cabinets.

Chuck

PS: Please remember that we are nearly 100% paperless and we would ask
that any reply to this message be by way of the "reply" feature on your
software, so that the entire string of correspondence can come back to us
and all be in one place.

===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: http://www.2spi.com
############################
==================================================




From MicroscopyL-request-at-ns.microscopy.com Sat Mar 5 11:50:51 2005



From: Jian-Guo Zheng :      j-zheng3-at-northwestern.edu
Date: Sat, 5 Mar 2005 11:57:59 -0600
Subject: [Microscopy] Pros and cons of in-column and post-column energy filter in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
We know some pros and cons of in-column and post-column energy filters
in TEM. To summarize them, I would like to hear from you who have
experiences with the filters.
Thanks, Jian



From MicroscopyL-request-at-ns.microscopy.com Sat Mar 5 12:05:08 2005



From: Ronald Anderson :      randerson20-at-tampabay.rr.com
Date: Sat, 05 Mar 2005 13:11:56 -0500
Subject: [Microscopy] Microscopy Today March 2005 Table of Contents

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Listers,

Here is the March 2005 Microscopy Today table of contents. I will close
the subscription list for this issue on Thursday March 10, 2005.

Microscopists in North America and MSA members anywhere can subscribe
for free. Anyone else may subscribe for US$35 per year (to PARTIALLY
cover postage). All subscriptions at http://www.microscopy-today.com
Thank you.


Turning Blood Cells into Oyster Shells
Stephen W. Carmichael, Mayo Clinic

New Developments in GEMINI® FESEM Technology
H. Jaksch, J-P Vermeulen, Carl Zeiss SMT Oberkochen, Germany

Crystal Scanner for Nano-Metrology Applications
Paul West, Zhiqiang Peng, Natalia Starostina, Pacific Nanotechnology, Inc.

Towards Optimal Imaging and Microanalysis in Variable Pressure and Low
Voltage SEM
Brendan J Griffin, The University of Western Australia

Effects of Abbe Condenser Spherical Aberration on Image Quality
Theodore M. Clarke, Metallurgical Failure Analysis Consultant

Experience with a Dicing Saw for Rapid Pre-FIB TEM Sample Preparation
Jim Conner*, James Beck*, and Bryan Tracy,** *Freescale Semiconductor,
Crolles, France and Austin, TX, **Spansion, Sunnyvale, CA

Digital Scanning, Archiving, and Transmitting Electron Micrographs
Joiner Cartwright, Jr., Baylor College of Medicine

What TEM Camera Should I Choose?
Michael Bode, Michael Wibbelt,* and Christoph Huelk*
Soft Imaging system Corp. *Soft Imaging System GmbH

There is Art in Science and Science in Art
Thomas H. Saunders, Amateur Microscopist

Sample Preparation for Textile Nanofiber Composites
R. Garcia, N. Fedorova, V. Knowlton, C. Oldham, B. Pourdeyhimi, NC State
University, Raleigh

Funding Opportunities for Acquiring Major Equipment from Federal
Granting Agencies M&M 2004 Core Facility Management – Part II: NSF
Organizer: Debby Sherman, Purdue University

New Resin for Repair of Bell Jar Chips
Owen P. Mills and Matt Huuki,* Michigan Tech. Univ., * Matt’s Auto
Glass, Houghton, MI

Inexpensive Digitization of an SEM
Henry C. Aldrich and Donna S. Williams, Univ. of Florida, Gainesville

Industry News

NetNotes



Ron Anderson, Editor
Microscopy Today



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 02:34:19 2005



From: Evren Cubukcu :      ecubukcu-at-hacettepe.edu.tr
Date: Mon, 07 Mar 2005 10:44:23 +0200
Subject: [Microscopy] RE: SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thank you very much for such valuable comments and suggestions.
I have been in SEM field for 2 years now and your contributions are really
amazing.

Evren




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 10:41:47 2005



From: Ian MacLaren :      i.maclaren-at-physics.gla.ac.uk
Date: Mon, 07 Mar 2005 16:50:28 +0000
Subject: [Microscopy] Re: Pros and cons of in-column and post-column energy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear all,
I'm sure there are better experts than me out there, but I just have a
few comments from my own experience.

Post column:
Pros:
can be attached to almost any TEM
Can be fairly easy to operate with proprietary software from manufacturer
Con:
Can have large post-column mag
- restricts field of view, cannot do low mag work (but see next two
points)
- partially fixed in newer models
- can be corrected by demagnifying image in projection lenses of
microscope, when this is set up correctly, easier on some microscopes
than others
Integration in/with microscope software - may vary with microscope
manufacturer - relies on good relations between the two manufacturers
Cannot use with film, must use CCD

In-column
Pros:
Can photograph image to film or CCD as you choose
What you see is what you get from fluorescent screen to film/CCD
No extra magnification - can do high or low mag, as you will (easier
then for energy filtered diffraction)
Should integrate perfectly into microscope software since microscope and
filter from same manufacturer
Cons:
Only available from some manufacturers (mainly LEO and JEOL)
Can be rather expensive
Limited choice of high voltages
Depends if you like the manufacturers proprietary software

Best wishes

--
Ian MacLaren
Department of Physics and Astronomy
University of Glasgow
Glasgow G12 8QQ
Scotland
http://www.ssp.gla.ac.uk/

****************************************************
FERROELECTRICS UK 2005: 26-27th April 2005
http://www.paisley.ac.uk/computing/ferroelectrics/
****************************************************



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 12:19:48 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 7 Mar 2005 10:44:09 -0800
Subject: [Microscopy] Re: Pros and cons of in-column and post-column energy filter in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Mar 5, 2005, at 9:57 AM, Jian-Guo Zheng wrote:

} We know some pros and cons of in-column and post-column energy filters
} in TEM. To summarize them, I would like to hear from you who have
} experiences with the filters.
}
Dear Jian,
We have a Gatan post-column imaging energy filter, which we use
exclusively for imaging of biological cryo-specimens. The images
obtained on this system are of very high quality. I have no experience
using the system for EELS or element mapping, nor do I have any
experience with in-column filters.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 12:31:02 2005



From: Mike Zemyan :      mzemyan-at-schaferlabs.com
Date: Mon, 7 Mar 2005 10:41:06 -0800
Subject: [Microscopy] RE: Secador Desiccator Cabinets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


In reading the description of these automatic dessicators, I notice that
they have 2 fans, one of which continually circulates air inside the
chamber. Does anyone know, are the fan bearings oil-lubricated? Is this
potentially a source of oil vapor contamination? We use dessicators (not
the automatic kind) to store cleaned vacuum parts, and so oil is a concern.

Michael Zemyan
Schafer Corporation

} -----Original Message-----
} From: Garber, Charles A. [mailto:cgarber-at-2spi.com]
} Sent: Saturday, March 05, 2005 8:11 AM
} To: MICROSCOPY BB
} Subject: [Microscopy] Secador Desiccator Cabinets
}
}
}
} --------------------------------------------------------------
} ----------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} --------------------------------------------------------------
} -----------------
}
} -- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --
}
} Michael Shaffer has written:
} ===================================================================
} Is anyone familiar with the automatic dessicating cabinets
} available ... e.g , "SECADOR" cabinets which are available
} from Electron Microscopy Sciences.
} How do they work? More importantly, how well do they work?
} ===================================================================
} The SecadorT family of desiccator cabinets are explained in
} detail (as well as their operation) on the SPI Supplies
} website at URL
} http://www.2spi.com/catalog/supp/desiccator-cabinets-secador.html
}
} To put it simply, the cabinets have an automatic desiccant
} regenerating station. It goes through an automatic cycle
} where by every so often (about twenty minutes), the
} regenerating station is closed off from the volume of the
} cabinet, the desiccant is heated, and the moisture is "blown
} off" to the outside. After the heat cycle, the doors to the
} outside are closed, and the doors to the inside of the
} chamber are opened, and the process of cabinet desiccation
} started all over again.
}
} SPI Supplies has sold these cabinets to a number of customers
} working not only in microscopy laboratories but into other
} settings where one would
} prefer to not have to handle particle-generating desiccants.
}
} How well do they work? They work as as well as any other
} desiccator cabinet but they work "better" if you factor in
} the fact that one does not have to rely on remembering to
} change desiccant to keep the contents of the chamber dry.
}
} Disclaimer: SPI Supplies is a distributor for the Secador
} line of desiccating cabinets.
}
} Chuck
}
} PS: Please remember that we are nearly 100% paperless and
} we would ask
} that any reply to this message be by way of the "reply"
} feature on your software, so that the entire string of
} correspondence can come back to us and all be in one place.
}
} ===================================================
} Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
} President
} SPI SUPPLIES FAX: 1-(610)-436-5755
} PO BOX 656 e-mail: cgarber-at-2spi.com
} West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com
}
}
} Look for us!
} ############################
} WWW: http://www.2spi.com
} ############################
} ==================================================
}
}



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 12:50:33 2005



From: john hoffpauir :      hoffpajo-at-yahoo.com
Date: Mon, 7 Mar 2005 10:59:45 -0800 (PST)
Subject: [Microscopy] Re: Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

i suspect those are intracelluar viruse particles.
more info on the tissue ie. is it a path speicimen
would be helpful.
john
--- Debby Sherman {dsherman-at-purdue.edu} wrote:
}
}
}
------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The
} Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help
}
http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
-------------------------------------------------------------------------------
}
} I'll take a stab at it and say it looks like virus
} in a crystalline array
} although size appears a bit small.
}
} Is there any signs of fibers in longitudinal
} orientation? If it is made up
} of cytoskeletal elements, i.e. fibers of different
} diameters, than I would
} expect to see some in longitudinal as well as
} x-sectional view.
}
} Debby
}
} Debby Sherman, Manager Phone:
} 765-494-6666
} Life Science Microscopy Facility FAX:
} 765-494-5896
} Purdue University E-mail:
} dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907
} http://www3.agriculture.purdue.edu/microscopy
}
}
} On 3/3/05 12:38 PM, "Aleksandr Mironov"
} {Aleksandr.Mironov-at-manchester.ac.uk}
} wrote:
}
} }
} }
} }
}
------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The
} Microscopy Society of America
} } To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} } On-Line Help
}
http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
}
------------------------------------------------------------------------------}
} -
} }
} } Dear collegues!
} }
} } Just recently I found the structure, which I
} cannot identify. It has
} } regular hexagonal pattern and localized in the
} cytosol of the cells in
} } embryonic tendons. It is not membrabous, some kind
} of cytoskeletal?
} }
} } Could you, please, take a look and hint what it
} could be?
} } The address for pictures is (structure is
} identified by arrows):
} } http://sysoj.spymac.net/image.jpg
} } http://sysoj.spymac.net/image2.jpg
} }
} } Regards,
}
}
}
}

__________________________________________________
Do You Yahoo!?
Tired of spam? Yahoo! Mail has the best spam protection around
http://mail.yahoo.com


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 14:25:31 2005



From: David Hall :      hall-at-aecom.yu.edu
Date: Mon, 7 Mar 2005 15:35:22 -0500
Subject: [Microscopy] Re: TEM Identifying intracellular structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I would vote for these to be an actin/myosin myofilament lattice.
The smaller actin fibrils surround each myosin rod in typical
fashion. The only surprising aspect is that the lattice is seen in
perfect cross-section and each separate portion of lattice seems to
be co-linear, suggesting that the whole cell is already polarized and
becoming anchored.

We've recently put up some images of nematode muscle that show these
features in comparison to vertebrate muscle: see Fig's 11, 15, 35
and 36 at the following two webpages:
www.wormatlas.org/handbook/mesodermal.htm/musclepartI.htm
www.wormatlas.org/handbook/mesodermal.htm/musclepartII.htm
--
David H. Hall, Ph.D.
Center for C. elegans Anatomy
Department of Neuroscience
Albert Einstein College of Medicine
1410 Pelham Parkway
Bronx, NY 10461

www.wormatlas.org
www.aecom.yu.edu/wormem

phone 718 430-2195
fax 718 430-2514


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 14:26:23 2005



From: Chaoying Ni :      cni-at-udel.edu
Date: Mon, 7 Mar 2005 15:41:02 -0500
Subject: [Microscopy] Re: Pros and cons of in-column and post-column

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Then Bill, in your case, GIF = CCD with a limited magnification range
and other inconveniences.

-Chao


-----Original Message-----
} From: Bill Tivol [mailto:tivol-at-caltech.edu]
Sent: Monday, March 07, 2005 1:44 PM
To: microscopy-at-msa.microscopy.com

On Mar 5, 2005, at 9:57 AM, Jian-Guo Zheng wrote:

} We know some pros and cons of in-column and post-column energy filters
} in TEM. To summarize them, I would like to hear from you who have
} experiences with the filters.
}
Dear Jian,
We have a Gatan post-column imaging energy filter, which we use
exclusively for imaging of biological cryo-specimens. The images
obtained on this system are of very high quality. I have no experience
using the system for EELS or element mapping, nor do I have any
experience with in-column filters.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 14:33:01 2005



From: Chaoying Ni :      cni-at-udel.edu
Date: Mon, 7 Mar 2005 15:47:03 -0500
Subject: [Microscopy] Re: Pros and cons of in-column and post-column

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Jian,

Ian made quite nice points. Is it also true that post-filter is better
for spectroscopy while in-column filter is better for energy-filtered
imaging? But I'm not so sure.
-Chao

-----Original Message-----
} From: Ian MacLaren [mailto:i.maclaren-at-physics.gla.ac.uk]
Sent: Monday, March 07, 2005 11:50 AM
To: Jian-Guo Zheng
Cc: Microscopy-at-microscopy.com

Dear all,
I'm sure there are better experts than me out there, but I just have a
few comments from my own experience.

Post column:
Pros:
can be attached to almost any TEM
Can be fairly easy to operate with proprietary software from
manufacturer
Con:
Can have large post-column mag
- restricts field of view, cannot do low mag work (but see next two
points)
- partially fixed in newer models
- can be corrected by demagnifying image in projection lenses of
microscope, when this is set up correctly, easier on some microscopes
than others
Integration in/with microscope software - may vary with microscope
manufacturer - relies on good relations between the two manufacturers
Cannot use with film, must use CCD

In-column
Pros:
Can photograph image to film or CCD as you choose
What you see is what you get from fluorescent screen to film/CCD
No extra magnification - can do high or low mag, as you will (easier
then for energy filtered diffraction)
Should integrate perfectly into microscope software since microscope and

filter from same manufacturer
Cons:
Only available from some manufacturers (mainly LEO and JEOL)
Can be rather expensive
Limited choice of high voltages
Depends if you like the manufacturers proprietary software

Best wishes

--
Ian MacLaren
Department of Physics and Astronomy
University of Glasgow
Glasgow G12 8QQ
Scotland
http://www.ssp.gla.ac.uk/

****************************************************
FERROELECTRICS UK 2005: 26-27th April 2005
http://www.paisley.ac.uk/computing/ferroelectrics/
****************************************************





From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 14:36:21 2005



From: Leslie Cummins :      gunther-at-aecom.yu.edu
Date: Mon, 07 Mar 2005 15:46:16 -0500
Subject: [Microscopy] Transfer of fungal cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Listers

We have a user that has some fungi attached to stone chips from Italian
monuments. He needs to get the fungi off the chips and onto a slide, so he
can do some fluorescence.

We have a few ideas to transfer the cells. First, to physically scrape
them off onto the slide. Second, transfer them using some sort of tape,
like plain old Scotch. But then, how do we get them to stick to the
slide. We are concerned that poly-lysine might not hold them down through
all his changes. How do we get them off the tape and onto the slide. How
do we try to keep the morphology as intact as possible?

If anyone has any ideas, we would love to hear your input.

Thanks in advance.
Leslie

Leslie Cummins www.aecom.yu.edu/aif
Analytical Imaging Facility
1300 Morris Park Ave
Forchheimer 639
Bronx, NY 10461
718-430-3547



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 14:39:34 2005



From: Becky Holdford :      r-holdford-at-ti.com
Date: Mon, 07 Mar 2005 14:48:46 -0600
Subject: [Microscopy] Re: Metal Corrosion Reference

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Anjeanette: this might be a good reference from ASM:
http://www.asminternational.org/Template.cfm?Section=BrowsebyTopic&template=Ecommerce/ProductDisplay.cfm&ProductID=10488
This is the ASM Handbook Volume 13: Corrosion.
Another good one might be "Corrosion: Understanding the Basics" from the
same place.

Anjeanette Ormonde wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Becky Holdford (r-holdford-at-ti.com)
972-995-2360
972-648-8743 (pager)
SC Packaging FA Development
Texas Instruments, Inc.
Dallas, TX
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 16:44:52 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 07 Mar 2005 14:54:33 -0800
Subject: [Microscopy] RE: SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


First of all, what method did you use to arrive
at the selection of an EVO SEM? Why not high vacuum?
I think that this needs to be factored if you're not sure.
What type of final specimen are you going to be analyzing?
If the soil is embedded, polished low mag imaged, you don't
necessarily need an ESEM. That said, if this is mostly all you
will be looking at, you can suffer the loss of resolution
of an ESEM. That you prefer not needing a chiller
is important. However, there are many advantages
in SEM systems that are only achieved by chiller-based
SEMs. You need to sort out all of the variables
for a SEM system first, I think, then select an
appropriate EDS system. In both cases, availability
of service resources is critical. If you don't have
a chiller, what cools the diffusion pump or turbo?
Some turbos can be air cooled.

If you are going to do an ESEM, it may likely
have a W filament or LaB6 at best. Thus, how
are you going to achieve 400K cps much less than
actually processing them? Furthermore, regardless
of the maximum cps stated, what filter time is this based on?
Short times can handle large cps at low resolution
and low fidelity. As filter time increases, dead time
increases and the existence of many counts becomes
irrellevant if not counterproductive. Most (my guess)
systems run about 2000 cps at 30% DT at 102uS time
constant. If you are concerned about resolution and
sensitivity, you should focus on obtaining long filter
time and { 35% DT. A system can certainly handle
hundreds of thousands cps but not likely at long filter
times. In this case, short filter times don't produce
good quant results. However, short times are valuable for
mapping. So, keep this in mind based on your application.
If you seek high resolution, you need a very good detector
and long filter times. Additionally, it seems out of
character to me that a W system would produce very high
counts compared to a LaB6 or especially an FESEM. But
I've only done EDS on FESEMs.

The other issue to keep in mind is that if you want to
do quant EDS, you will need an EDS package that allows
for high pressure quant. I use EDAX Genesis VIP Quant
for variable pressure SEM (usually 5-50Pa). It is selectable
for VP or high vacuum. Furthermore, it provides for
ZAF, PhiZAF and PhiRhoZAF quant in VP or HV modes.
The user interface is very nice, IMO. I have used the
Rontec as well and got good results in a high vacuum
system. I don't know if they can handle high pressure
situations.

There have been discussions about standardless quant
on this list before. The method I use is what EDAX calls
SEC--a standards correction factor. This is where the
quant value is typically value*SEC where SEC=1. However,
the SEC value can be changed to a value that makes the
quant result match a standard. Thus, quant results are
done routinely in a standardless manner but are traced
back to a standard. This SEC check should be done on
a routine basis (I do it every quarter). The other thing
to check is that the peak for each detected element lines
up with its respective eV for whichever shell is being
examined. This value can shift slightly. Sometimes,
an element's eV value might be high or low by a small
amount and confuse a convolution (peak pileup).

As far as the detector being LN2-free, I agree very much.
The tradeoffs for this option are cost and performance.
Without getting at or close to LN2 temperature, detector
performance will suffer. I use the EDAX Cryospec LN2-free
detector. This detector produces the same performance
as an LN2 detector. Currently, I am getting 131.5eV
resolution and can detect down to and including Be.
But of course, YMMV.

gary g.

At 12:44 AM 3/7/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 17:06:07 2005



From: Tamara Howard :      thoward-at-unm.edu
Date: Mon, 7 Mar 2005 16:15:54 -0700 (MST)
Subject: [Microscopy] Re: Transfer of fungal cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Double-sided scotch tape?

On Mon, 7 Mar 2005, Leslie Cummins wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Hello Listers
}
} We have a user that has some fungi attached to stone chips from Italian
} monuments. He needs to get the fungi off the chips and onto a slide, so he
} can do some fluorescence.
}
} We have a few ideas to transfer the cells. First, to physically scrape them
} off onto the slide. Second, transfer them using some sort of tape, like
} plain old Scotch. But then, how do we get them to stick to the slide. We
} are concerned that poly-lysine might not hold them down through all his
} changes. How do we get them off the tape and onto the slide. How do we try
} to keep the morphology as intact as possible?
}
} If anyone has any ideas, we would love to hear your input.
}
} Thanks in advance.
} Leslie
}
} Leslie Cummins www.aecom.yu.edu/aif
} Analytical Imaging Facility
} 1300 Morris Park Ave
} Forchheimer 639
} Bronx, NY 10461
} 718-430-3547
}
}

|--------------------------------------------------|
Tamara Howard
Department of Cell Biology and Physiology
University of New Mexico - Health Sciences Center
Albuquerque, NM 87131
thoward-at-unm.edu
|--------------------------------------------------|


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 17:28:42 2005



From: kak-at-medicine.wisc.edu (by way of Ask-A-Microscopist)
Date: Mon, 7 Mar 2005 17:39:09 -0600
Subject: [Microscopy] viaWWW: binocular compound microscope as a hobbyist

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (kak-at-medicine.wisc.edu) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Monday, March 7, 2005 at 15:52:14
---------------------------------------------------------------------------

Email: kak-at-medicine.wisc.edu
Name: Kip Kircher

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I am researching the purchase of a binocular compound microscope as a hobbyist. The LOMO BMH4-BF seems the best deal at $445. I find it ugly however and am suspicious of the quality. Is this the best under $500 deal available?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 17:51:48 2005



From: Angela V. Klaus :      avklaus-at-amnh.org
Date: Mon, 7 Mar 2005 19:00:29 -0500 (EST)
Subject: [Microscopy] LN2-free x-ray detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Gary,

I'm thinking about switching to an LN2-free EDS detector. It sounds like
you're fairly happy with yours, but I was wondering what "YMMV" stands for
in the below portion of your note:

} As far as the detector being LN2-free, I agree very much.
} The tradeoffs for this option are cost and performance.
} Without getting at or close to LN2 temperature, detector
} performance will suffer. I use the EDAX Cryospec LN2-free
} detector. This detector produces the same performance
} as an LN2 detector. Currently, I am getting 131.5eV
} resolution and can detect down to and including Be.
} But of course, YMMV.

I'd be really interested to hear what other experiences people have had
with liquid nitrogen-free EDS detectors.

All best,

Angela

--
Angela V. Klaus, Ph.D.
Director, Microscopy and Imaging Facility
American Museum of Natural History
Central Park West and 79th Street
New York, NY 10024 USA
Email: avklaus-at-amnh.org
Tel: 212-796-5977
Fax: 212-496-3480

}
}
} ------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America








From MicroscopyL-request-at-ns.microscopy.com Mon Mar 7 20:27:19 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 07 Mar 2005 18:37:32 -0800
Subject: [Microscopy] Re: LN2-free x-ray detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Reply made off-list.

gary g.



At 04:00 PM 3/7/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 09:20:35 2005



From: Peter Van Osta :      pvosta-at-maia-scientific.com
Date: Tue, 08 Mar 2005 16:30:02 +0100
Subject: [Microscopy] Filter for DRAQ5 and GFP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,

Which filterset (Omega, ...) and dichroic is mostly used for dual label
experiments with the nuclear probe DRAQ5 and a GFP label ?

Regards,

Peter Van Osta

----------------------------------------------
Peter Van Osta, MD

Director Imaging
MAIA SCIENTIFIC
Cipalstraat 3
B-2440 Geel, Belgium
Tel.: +32 (0)14 570 620
Mobile: +32 (0)497 228 725
Fax.: +32 (0)14 570 621
Email: pvosta-at-maia-scientific.com
Website: www.maia-scientific.com
A Harvard Bioscience Company
----------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 10:19:14 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Tue, 08 Mar 2005 08:28:42 -0800
Subject: [Microscopy] Re: LN2-free x-ray detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Well, I suppose for hassle reduction purposes.
It seems that whenever products are discussed
rather than procedures, the thread can turn into
a contrarian epic. I'm just trying to answer the
question based on what I have, without speculation
about what something else might do.

Here is the response:

/# begin #/
I'm fairly happy with mine. This is the third
unit I have had of the Cryospec. The system
performs extraordinarily well when all is stable
and nice. What the system hates is being turned off
for awhile and then turned back on. In this case,
it may not achieve low enough temperature for the
HV to turn on--dead system.

The Cryospec uses essentially the same basic Si(Li)
Moxtec detector as any EDAX (or other) LN2 detector.
The difference is that there is a cryo pump which
is a separate box with light weight hoses to a
screwed-on unit for the detector that cools the
detector. The cryo pump is made by MMR and does a
very good job when stable. It is consistent and
dependable. The resolution of the Cryospec detector
is excellent and does this without LN2. It just sits
and runs. I really like the EDAX Genesis software
package. Easy to use and their HPD feature is great
for identifying wrong element peaks.

If you get the mapping option and image collection
option, you will then have a digital image capture
SEM. Plus, you can do maps based on detected Zs or
whatever you want. And there are many other features
of the Genesis package. I also have a remote for my
separate PC and notebook PC so I can analyze EDS
spectra separate from the SEM system. Very handy.
/# end #/

The cooling versus non-cooling was intended to
address SEM units only. Cooled versus non-cooled can
have a big impact on what is sitting in your room (foot print)
and what image results and performance you obtain.
Specs are one thing but actually achieving them is
another. An un-coated specimen imaged at 100V, 4mm WD
and 150KX would, I think, be rather difficult without
a cooled SEM. And the design of the SEM would
have to be quite different from traditional designs.

Zeiss achieves this in high vacuum mode on Supra and
Ultra FESEMs. I don't know about their EVSEM performance.
That was the crux of the suggestion of matching a SEM to
requirements rather than stating the SEM type up front.

gary g.


At 05:18 AM 3/8/2005, you wrote:
} Gary,
}
} Just curious as to why you chose not to share the reply. I am also
} interested in learning pros and cons about N2-free detectors.
} In your E-mail below you seem to indicate that not cooling may have negative
} affects on performance as well as increased costs. Then you state that your
} N2-free detector works equivalent to those requiring cooling.
}
} Could you explain this apparent contradiction?
}
} Debby
}
} Debby Sherman, Manager Phone: 765-494-6666
} Life Science Microscopy Facility FAX: 765-494-5896
} Purdue University E-mail: dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907
} http://www3.agriculture.purdue.edu/microscopy
}
} On 3/7/05 9:37 PM, "Gary Gaugler" {gary-at-gaugler.com} wrote:
}
} }
} }
} }
} ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
} ------------------------------------------------------------------------------}
} -
} }
} } Reply made off-list.
} }
} } gary g.
} }
} }
} }
} } At 04:00 PM 3/7/2005, you wrote:
} }
} }
} } }
} -----------------------------------------------------------------------------} }
} -
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} } } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} -----------------------------------------------------------------------------
} } } --
} } }
} } } Hi Gary,
} } }
} } } I'm thinking about switching to an LN2-free EDS detector. It sounds like
} } } you're fairly happy with yours, but I was wondering what "YMMV" stands for
} } } in the below portion of your note:
} } }
} } } } As far as the detector being LN2-free, I agree very much.
} } } } The tradeoffs for this option are cost and performance.
} } } } Without getting at or close to LN2 temperature, detector
} } } } performance will suffer. I use the EDAX Cryospec LN2-free
} } } } detector. This detector produces the same performance
} } } } as an LN2 detector. Currently, I am getting 131.5eV
} } } } resolution and can detect down to and including Be.
} } } } But of course, YMMV.
} } }
} } } I'd be really interested to hear what other experiences people have had
} } } with liquid nitrogen-free EDS detectors.
} } }
} } } All best,
} } }
} } } Angela
} } }
} } } --
} } } Angela V. Klaus, Ph.D.
} } } Director, Microscopy and Imaging Facility
} } } American Museum of Natural History
} } } Central Park West and 79th Street
} } } New York, NY 10024 USA
} } } Email: avklaus-at-amnh.org
} } } Tel: 212-796-5977
} } } Fax: 212-496-3480
} } }
} } } }
} } } }
} } } }
} } }
} -----------------------------------------------------------------------------} }
} -
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe --
} } } } http://www.microscopy.com/MicroscopyListserver
} } } } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} -----------------------------------------------------------------------------
} } } --
} } } }
} } } }
} } } } First of all, what method did you use to arrive
} } } } at the selection of an EVO SEM? Why not high vacuum?
} } } } I think that this needs to be factored if you're not sure.
} } } } What type of final specimen are you going to be analyzing?
} } } } If the soil is embedded, polished low mag imaged, you don't
} } } } necessarily need an ESEM. That said, if this is mostly all you
} } } } will be looking at, you can suffer the loss of resolution
} } } } of an ESEM. That you prefer not needing a chiller
} } } } is important. However, there are many advantages
} } } } in SEM systems that are only achieved by chiller-based
} } } } SEMs. You need to sort out all of the variables
} } } } for a SEM system first, I think, then select an
} } } } appropriate EDS system. In both cases, availability
} } } } of service resources is critical. If you don't have
} } } } a chiller, what cools the diffusion pump or turbo?
} } } } Some turbos can be air cooled.
} } } }
} } } } If you are going to do an ESEM, it may likely
} } } } have a W filament or LaB6 at best. Thus, how
} } } } are you going to achieve 400K cps much less than
} } } } actually processing them? Furthermore, regardless
} } } } of the maximum cps stated, what filter time is this based on?
} } } } Short times can handle large cps at low resolution
} } } } and low fidelity. As filter time increases, dead time
} } } } increases and the existence of many counts becomes
} } } } irrellevant if not counterproductive. Most (my guess)
} } } } systems run about 2000 cps at 30% DT at 102uS time
} } } } constant. If you are concerned about resolution and
} } } } sensitivity, you should focus on obtaining long filter
} } } } time and { 35% DT. A system can certainly handle
} } } } hundreds of thousands cps but not likely at long filter
} } } } times. In this case, short filter times don't produce
} } } } good quant results. However, short times are valuable for
} } } } mapping. So, keep this in mind based on your application.
} } } } If you seek high resolution, you need a very good detector
} } } } and long filter times. Additionally, it seems out of
} } } } character to me that a W system would produce very high
} } } } counts compared to a LaB6 or especially an FESEM. But
} } } } I've only done EDS on FESEMs.
} } } }
} } } } The other issue to keep in mind is that if you want to
} } } } do quant EDS, you will need an EDS package that allows
} } } } for high pressure quant. I use EDAX Genesis VIP Quant
} } } } for variable pressure SEM (usually 5-50Pa). It is selectable
} } } } for VP or high vacuum. Furthermore, it provides for
} } } } ZAF, PhiZAF and PhiRhoZAF quant in VP or HV modes.
} } } } The user interface is very nice, IMO. I have used the
} } } } Rontec as well and got good results in a high vacuum
} } } } system. I don't know if they can handle high pressure
} } } } situations.
} } } }
} } } } There have been discussions about standardless quant
} } } } on this list before. The method I use is what EDAX calls
} } } } SEC--a standards correction factor. This is where the
} } } } quant value is typically value*SEC where SEC=1. However,
} } } } the SEC value can be changed to a value that makes the
} } } } quant result match a standard. Thus, quant results are
} } } } done routinely in a standardless manner but are traced
} } } } back to a standard. This SEC check should be done on
} } } } a routine basis (I do it every quarter). The other thing
} } } } to check is that the peak for each detected element lines
} } } } up with its respective eV for whichever shell is being
} } } } examined. This value can shift slightly. Sometimes,
} } } } an element's eV value might be high or low by a small
} } } } amount and confuse a convolution (peak pileup).
} } } }
} } } } As far as the detector being LN2-free, I agree very much.
} } } } The tradeoffs for this option are cost and performance.
} } } } Without getting at or close to LN2 temperature, detector
} } } } performance will suffer. I use the EDAX Cryospec LN2-free
} } } } detector. This detector produces the same performance
} } } } as an LN2 detector. Currently, I am getting 131.5eV
} } } } resolution and can detect down to and including Be.
} } } } But of course, YMMV.
} } } }
} } } } gary g.
} } } }
} } } } At 12:44 AM 3/7/2005, you wrote:
} } } }
} } } }
} } } } } ------------------------------------------------------------------------
} } } ------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe --
} } } } } http://www.microscopy.com/MicroscopyListserver
} } } } } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} -----------------------------------------------------------------------------
} } } --
} } } Thank you very much for such valuable comments and suggestions.
} } } } } I have been in SEM field for 2 years now and your contributions are
} } } } } really
} } } } } amazing.
} } } } } Evren
} } } }
} } } }
} } } }
} }
} }



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 15:19:17 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Tue, 08 Mar 2005 13:28:55 -0800
Subject: [Microscopy] Re: RE: SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thanks. Yes. The point is that being able to handle
many counts only is relevant if the system will produce
them. With a W emitter, one would need high KV, big
aperture and relaxed condenser to get high counts.
Then, the result is poor resolution. My FESEM will
do up to at least 60K cps depending on KV, gun aperture
size, high or low current mode and Z of specimen. but
at high cps, in order to keep DT { 35%, filter time
is short. I'd rather have good resolution and just
take a little longer to get good quant data.

The EDAX Genesis has what they call VIP Quant for doing
variable pressure quant. It does like you guessed. One
takes a reading at one pressure then another at a different
pressure. It then extrapolates and produces the same type
of output as it does at high vacuum (HV or VP is selected
by a click box in the app).

when quant is produced, there is quant data along with
a separate listing of intensities for each element and
the intensity ratio error. If the ratio error is less
than about 10% or so, the result is considered OK. If
around 20% or greater, the result is bad--likely, wrong
element was tagged. So, go back and find out what the
correct element is close to the bad Z. this usually
does not happen since the HPD (halographic peak deconvolution)
shows if the user correctly identified elements. It also
shows if the user missed an element in a pile up of Zs.

This all works in high vacuum and VP modes. I'm running
between 20-50Pa...just enough to get rid of charging
with insulating specimens. Each one is slightly different.
If I recall correctly, VIP Quant also works for EVSEM (higher pressures),
not sure about this.

The way I do phase very nicely is with EDS coupled with EBSD.
Even so, with just EDS, Genesis will do whole scan measurements
or spot or ROI (option). This way, find an area and analyze it.
Then, do quant. If you want to know where stuff is, do a map
based on the collected Z list. But the best way is EDS+EBSD.

gary g.



At 10:13 AM 3/8/2005, you wrote:
} Nice comments about resolution and count rate. I don't think I've ever
} gotten more than 20 kcps out of our tungsten gun. FE would certainly make
} a difference.
}
} I am curious about your comments about high-pressure quant. Does EDAX have
} a special package for that? What does it do?
}
} I know that we get contributions from the neighboring area and have to be
} very careful when trying to do quant at high pressures (20-100 Pa). I was
} looking at gold-bearing minerals for a fellow last week. The gangue
} mineral was silicious so everything had silicon in it. Likewise, I get
} bogus Al-Cu analyses when I have a small Cu-rich phase dispersed in an
} Al-rich matrix. I can't quite imagine how they would handle the problem
} without a lot of knowledge about the specimen. Do they perform analyses
} under multiple pressures and extrapolate back to zero pressure?
}
} Curious Warren
}
} At 04:54 PM 03/07/05, you wrote:
}
} } First of all, what method did you use to arrive
} } at the selection of an EVO SEM? Why not high vacuum?
} } I think that this needs to be factored if you're not sure.
} } What type of final specimen are you going to be analyzing?
} } If the soil is embedded, polished low mag imaged, you don't
} } necessarily need an ESEM. That said, if this is mostly all you
} } will be looking at, you can suffer the loss of resolution
} } of an ESEM. That you prefer not needing a chiller
} } is important. However, there are many advantages
} } in SEM systems that are only achieved by chiller-based
} } SEMs. You need to sort out all of the variables
} } for a SEM system first, I think, then select an
} } appropriate EDS system. In both cases, availability
} } of service resources is critical. If you don't have
} } a chiller, what cools the diffusion pump or turbo?
} } Some turbos can be air cooled.
} }
} } If you are going to do an ESEM, it may likely
} } have a W filament or LaB6 at best. Thus, how
} } are you going to achieve 400K cps much less than
} } actually processing them? Furthermore, regardless
} } of the maximum cps stated, what filter time is this based on?
} } Short times can handle large cps at low resolution
} } and low fidelity. As filter time increases, dead time
} } increases and the existence of many counts becomes
} } irrellevant if not counterproductive. Most (my guess)
} } systems run about 2000 cps at 30% DT at 102uS time
} } constant. If you are concerned about resolution and
} } sensitivity, you should focus on obtaining long filter
} } time and { 35% DT. A system can certainly handle
} } hundreds of thousands cps but not likely at long filter
} } times. In this case, short filter times don't produce
} } good quant results. However, short times are valuable for
} } mapping. So, keep this in mind based on your application.
} } If you seek high resolution, you need a very good detector
} } and long filter times. Additionally, it seems out of
} } character to me that a W system would produce very high
} } counts compared to a LaB6 or especially an FESEM. But
} } I've only done EDS on FESEMs.
} }
} } The other issue to keep in mind is that if you want to
} } do quant EDS, you will need an EDS package that allows
} } for high pressure quant. I use EDAX Genesis VIP Quant
} } for variable pressure SEM (usually 5-50Pa). It is selectable
} } for VP or high vacuum. Furthermore, it provides for
} } ZAF, PhiZAF and PhiRhoZAF quant in VP or HV modes.
} } The user interface is very nice, IMO. I have used the
} } Rontec as well and got good results in a high vacuum
} } system. I don't know if they can handle high pressure
} } situations.
} }
} } There have been discussions about standardless quant
} } on this list before. The method I use is what EDAX calls
} } SEC--a standards correction factor. This is where the
} } quant value is typically value*SEC where SEC=1. However,
} } the SEC value can be changed to a value that makes the
} } quant result match a standard. Thus, quant results are
} } done routinely in a standardless manner but are traced
} } back to a standard. This SEC check should be done on
} } a routine basis (I do it every quarter). The other thing
} } to check is that the peak for each detected element lines
} } up with its respective eV for whichever shell is being
} } examined. This value can shift slightly. Sometimes,
} } an element's eV value might be high or low by a small
} } amount and confuse a convolution (peak pileup).
} }
} } As far as the detector being LN2-free, I agree very much.
} } The tradeoffs for this option are cost and performance.
} } Without getting at or close to LN2 temperature, detector
} } performance will suffer. I use the EDAX Cryospec LN2-free
} } detector. This detector produces the same performance
} } as an LN2 detector. Currently, I am getting 131.5eV
} } resolution and can detect down to and including Be.
} } But of course, YMMV.
} }
} } gary g.
}



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 16:17:29 2005



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Tue, 08 Mar 2005 17:27:43 -0500
Subject: [Microscopy] Spelling question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Is there any opinion as to the preferred way of spelling:

Bremmstrahlung (948 hits)

or

Bremsstrahlung (317,000 hits)

radiation?

If one does Google searches on each of the two spelllings, there is indeed
indicated a preference of one over the other, as indicated above. Does that
mean the second spelling is the correct one?

Chuck


============================================
Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================





From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 18:06:09 2005



From: Mike Bode :      Mike.Bode-at-soft-imaging.net
Date: Tue, 8 Mar 2005 17:14:04 -0700
Subject: [Microscopy] Spelling question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Chuck,

the word is German and comes from "bremsen" (to break, slow down) and
"Strahlung" (radiation). The correct spelling would be "Bremsstrahlung".

mike


Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: Garber, Charles A. [mailto:cgarber-at-2spi.com]
Sent: Tuesday, March 08, 2005 15:28
To: MICROSCOPY BB

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Is there any opinion as to the preferred way of spelling:

Bremmstrahlung (948 hits)

or

Bremsstrahlung (317,000 hits)

radiation?

If one does Google searches on each of the two spelllings, there is indeed
indicated a preference of one over the other, as indicated above. Does that
mean the second spelling is the correct one?

Chuck


============================================
Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.2spi.com
########################
============================================






From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 18:25:35 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Tue, 8 Mar 2005 16:50:03 -0800
Subject: [Microscopy] Re: Spelling question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Mar 8, 2005, at 2:27 PM, Garber, Charles A. wrote:

} Bremsstrahlung (317,000 hits)
}
Dear Chuck,
This spelling is correct; my limited memory of German is that "brems"
is from "braking", "strahl" means "radiate" or "stream" (the verb), and
the suffix "ung" makes the word a noun (like "ing"). Those on the list
who know more German than I can correct me.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 18:33:19 2005



From: Becky Holdford :      r-holdford-at-ti.com
Date: Tue, 08 Mar 2005 18:43:29 -0600
Subject: [Microscopy] Texas Society for Microscopy 40th Anniversary meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I would like to take this opportunity to invite you to the 40th anniversary
celebration of the Texas Society for Microscopy. The spring meeting will
be held April 14-16 in Las Colinas, TX at the Dallas Marriott and marks a
milestone for our society. Below are some of the highlights for this meeting:

+ Parallel sessions for the biological sciences and materials sciences
+ Special Friday afternoon session to celebrate our history
+ Workshop " Low Voltage Variable Pressure SEM for Nanotechnology"
presented by Steve Joens at Hitachi
+ Guest speakers
- "Failure Analysis of Non Visual Fails on Integrated Circuits
Using Nano Probing in an SEM" presented by Dr. Richard Stallcup III, Zyvex
Corporation
- "Optical Coherence Microscopy, A Technology for Rapid, In Vivo,
Nondestructive Visualization of Plants and Plants Cells" presented by Dr.
June Medford, Biology Dept, Colorado State University
+ Vendor exhibits
+ Platform presentations and poster session
+ Photo contest

Additional information can be found at http://www.texasmicroscopy.org/

Hope to see you in April.

Regards,
Texas Society for Microscopy



--
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Becky Holdford (r-holdford-at-ti.com)
972-995-2360
972-648-8743 (pager)
SC Packaging FA Development
Texas Instruments, Inc.
Dallas, TX
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 18:55:21 2005



From: Ken Converse :      kenconverse-at-qualityimages.biz
Date: Tue, 8 Mar 2005 20:04:58 -0500
Subject: [Microscopy] RE: SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Gary,
I believe the countrates are correct. If you check
http://www.rontecusa.com/products.htm
I think you'll find they have a new detector design. It's lN2 free and
operates very fast. The resolution is not quite as good as their lN2 cooled
detectors, but the speeds are real.

I have no financial interest but did hear a talk on them in Woods Hole, MA
with NESM last May.

Ken Converse

owner 
QUALITY IMAGES
Servicing Scanning Electron Microscopes
Since 1981
16 Creek Rd.
Delta, PA 17314
717-456-5491
Fax 717-456-7996
kenconverse-at-qualityimages.biz
qualityimages.biz


-----Original Message-----
} From: Gary Gaugler [mailto:gary-at-gaugler.com]
Sent: Monday, March 07, 2005 5:55 PM
To: Evren Cubukcu
Cc: MSA listserver


First of all, what method did you use to arrive
at the selection of an EVO SEM? Why not high vacuum?
I think that this needs to be factored if you're not sure.
What type of final specimen are you going to be analyzing?
If the soil is embedded, polished low mag imaged, you don't
necessarily need an ESEM. That said, if this is mostly all you
will be looking at, you can suffer the loss of resolution
of an ESEM. That you prefer not needing a chiller
is important. However, there are many advantages
in SEM systems that are only achieved by chiller-based
SEMs. You need to sort out all of the variables
for a SEM system first, I think, then select an
appropriate EDS system. In both cases, availability
of service resources is critical. If you don't have
a chiller, what cools the diffusion pump or turbo?
Some turbos can be air cooled.

If you are going to do an ESEM, it may likely
have a W filament or LaB6 at best. Thus, how
are you going to achieve 400K cps much less than
actually processing them? Furthermore, regardless
of the maximum cps stated, what filter time is this based on?
Short times can handle large cps at low resolution
and low fidelity. As filter time increases, dead time
increases and the existence of many counts becomes
irrellevant if not counterproductive. Most (my guess)
systems run about 2000 cps at 30% DT at 102uS time
constant. If you are concerned about resolution and
sensitivity, you should focus on obtaining long filter
time and { 35% DT. A system can certainly handle
hundreds of thousands cps but not likely at long filter
times. In this case, short filter times don't produce
good quant results. However, short times are valuable for
mapping. So, keep this in mind based on your application.
If you seek high resolution, you need a very good detector
and long filter times. Additionally, it seems out of
character to me that a W system would produce very high
counts compared to a LaB6 or especially an FESEM. But
I've only done EDS on FESEMs.

The other issue to keep in mind is that if you want to
do quant EDS, you will need an EDS package that allows
for high pressure quant. I use EDAX Genesis VIP Quant
for variable pressure SEM (usually 5-50Pa). It is selectable
for VP or high vacuum. Furthermore, it provides for
ZAF, PhiZAF and PhiRhoZAF quant in VP or HV modes.
The user interface is very nice, IMO. I have used the
Rontec as well and got good results in a high vacuum
system. I don't know if they can handle high pressure
situations.

There have been discussions about standardless quant
on this list before. The method I use is what EDAX calls
SEC--a standards correction factor. This is where the
quant value is typically value*SEC where SEC=1. However,
the SEC value can be changed to a value that makes the
quant result match a standard. Thus, quant results are
done routinely in a standardless manner but are traced
back to a standard. This SEC check should be done on
a routine basis (I do it every quarter). The other thing
to check is that the peak for each detected element lines
up with its respective eV for whichever shell is being
examined. This value can shift slightly. Sometimes,
an element's eV value might be high or low by a small
amount and confuse a convolution (peak pileup).

As far as the detector being LN2-free, I agree very much.
The tradeoffs for this option are cost and performance.
Without getting at or close to LN2 temperature, detector
performance will suffer. I use the EDAX Cryospec LN2-free
detector. This detector produces the same performance
as an LN2 detector. Currently, I am getting 131.5eV
resolution and can detect down to and including Be.
But of course, YMMV.

gary g.

At 12:44 AM 3/7/2005, you wrote:


} ---------------------------------------------------------------------------
---
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America








From MicroscopyL-request-at-ns.microscopy.com Tue Mar 8 19:08:05 2005



From: Ellery Frahm :      frah0010-at-tc.umn.edu
Date: Tue, 8 Mar 2005 19:18:09 -0600
Subject: [Microscopy] Re: Spelling question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Based on the German words for "braking" (bremsen) and "radiation"
(strahlung), the spelling "bremsstrahlung" seems correct, as the
internet masses would seem to suggest. I also cannot find the spelling
"bremmstrahlung" in the books I have at hand. I would suggest sticking
with "bremsstrahlung" as the proper spelling. However, if any German
speakers care to differ with me, I'll happily stand corrected.

Best,
Ellery

--------------------
Ellery E. Frahm
Research Fellow & Manager
Electron Microprobe Laboratory
University of Minnesota - Twin Cities
Department of Geology & Geophysics
Lab Website: http://probelab.geo.umn.edu



On Mar 8, 2005, at 4:27 PM, Garber, Charles A. wrote:

} Is there any opinion as to the preferred way of spelling:
}
} Bremmstrahlung (948 hits)
}
} or
}
} Bremsstrahlung (317,000 hits)
}
} radiation?
}
} If one does Google searches on each of the two spelllings, there is
} indeed
} indicated a preference of one over the other, as indicated above.
} Does that
} mean the second spelling is the correct one?
}
} Chuck
}
} ============================================
} Charles A. Garber, Ph. D. Ph: 1-610-436-5400
} President 1-800-2424-SPI
} SPI SUPPLIES FAX: 1-610-436-5755
} PO BOX 656 e-mail:cgarber-at-2spi.com
} West Chester, PA 19381-0656 USA




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 00:59:09 2005



From: randy-nessler-at-uiowa.edu (by way of MicroscopyListserver)
Date: Wed, 9 Mar 2005 01:09:12 -0600
Subject: [Microscopy] viaWWW: Insurance Service Provider Vs OEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (randy-nessler-at-uiowa.edu) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
Tuesday, March 8, 2005 at 15:32:26
---------------------------------------------------------------------------

Email: randy-nessler-at-uiowa.edu
Name: Randy Nessler

Organization: University of Iowa

Title-Subject: [Microscopy] [Filtered] MListserver:Insurance Service Provider Vs OEM

Question: Here we go again. I realize that this has been discussed
here before but we are faced with this decision again. I have been
lead to believe that a Big Ten Consortium has been formed between an
insurance provider and the Universities that are members of the Big
Ten. Our purchasing department has been involved in our discussions
with the insurance company (to the point of being a proponent?). I
would like to hear from any Big Ten University EM labs (or any
EM/confocal labs for that matter) that have evaluated this service
contract option recently. Anybody out there????
Thanks,
Randy Nessler
Phone 319-335-8142
randy-nessler-at-uiowa.edu

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 00:59:31 2005



From: mitrah-at-uci.edu (by way of MicroscopyListserver)
Date: Wed, 9 Mar 2005 01:09:39 -0600
Subject: [Microscopy] viaWWW: Myelin Fixation for EM Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (mitrah-at-uci.edu) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
Tuesday, March 8, 2005 at 17:46:59
---------------------------------------------------------------------------

Email: mitrah-at-uci.edu
Name: Mitra Hooshmand

Organization: UC Irvine

Title-Subject: [Microscopy] [Filtered] MListserver: Myelin Fixation for EM Analysis

Question: I have been unsuccessfully attempting at obtaining optimal
myelin morphology for EM analysis. My ultimate goal is to do
immuno-EM on the spinal cord. However, I have not been able to get
myelin ultrastructure preservation even in doing EM alone.

If you have any recommendations/suggestions or a protocol that may
guide me, I would be most grateful. I am on an unforgiving deadline
and in dire need for some guidance. Thank you, immensely, in advance.

Mitra Hooshmand
Graduate Student Researcher
UC Irvine
mitrah-at-uci.edu

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 03:08:01 2005



From: Reinhard Rachel :      reinhard.rachel-at-biologie.uni-regensburg.de
Date: Wed, 09 Mar 2005 10:17:23 +0100
Subject: [Microscopy] Spelling question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Concerning the spelling and meaning of "Bremsstrahlung":

the comments given by Mike Bode and Bill Tivol are absolutely correct.

kind regards from Germany,

Reinhard

-------------------------------
PD Dr.Reinhard Rachel
Universität Regensburg
Lehrstuhl für Mikrobiologie
Universitaetsstr. 31
D-93053 Regensburg
tel.: +49 941 943 4534
fax.: +49 941 943 1824




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 03:44:57 2005



From: Gerd Leitinger :      gerd.leitinger-at-meduni-graz.at
Date: Wed, 09 Mar 2005 10:55:56 +0100
Subject: [Microscopy] Re: Re: Spelling question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Von: Ellery Frahm {frah0010-at-tc.umn.edu}
Betreff: Re: Spelling question
Datum: Tue, 8 Mar 2005 19:18:09 -0600
An: MICROSCOPY BB
{Microscopy-at-MSA.Microscopy.com}


"Bremsstrahlung" is the correct spelling.
As others have mentioned, it is derived from "bremsen (braking)" and
"Strahlung (radiation)".

The first s is pronounced like an "s", the second one like a "sh".
We have recently had a spelling reform wich changed the spelling of
several words, especially words containing sharp "ß"'s and double
"ss"`s, but to my knowledge the word "bremsen" has not been affected.

best wishes

Gerd

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe --
http://www.microscopy.com/MicroscopyListserver
} On-Line Help
http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Based on the German words for "braking" (bremsen) and "radiation"
} (strahlung), the spelling "bremsstrahlung" seems correct, as the
} internet masses would seem to suggest. I also cannot find the
spelling
} "bremmstrahlung" in the books I have at hand. I would suggest
sticking
} with "bremsstrahlung" as the proper spelling. However, if any
German
} speakers care to differ with me, I'll happily stand corrected.
}
} Best,
} Ellery
}
} --------------------
} Ellery E. Frahm
} Research Fellow & Manager
} Electron Microprobe Laboratory
} University of Minnesota - Twin Cities
} Department of Geology & Geophysics
} Lab Website: http://probelab.geo.umn.edu
}
}
}
} On Mar 8, 2005, at 4:27 PM, Garber, Charles A. wrote:
}
} } Is there any opinion as to the preferred way of spelling:
} }
} } Bremmstrahlung (948 hits)
} }
} } or
} }
} } Bremsstrahlung (317,000 hits)
} }
} } radiation?
} }
} } If one does Google searches on each of the two spelllings, there is
} } indeed
} } indicated a preference of one over the other, as indicated above.
} } Does that
} } mean the second spelling is the correct one?
} }
} } Chuck
} }
} } ============================================
} } Charles A. Garber, Ph. D. Ph: 1-610-436-5400
} } President 1-800-2424-SPI
} } SPI SUPPLIES FAX: 1-610-436-5755
} } PO BOX 656 e-mail:cgarber-at-2spi.com
} } West Chester, PA 19381-0656 USA
}
}
}




Dr. Gerd Leitinger

Institut für Zellbiologie, Histologie und Embryologie
Medizinische Universität Graz
Harrachgasse 21
A-8010 Graz
Austria

Tel. ++43 316 380 4237
Fax. ++43 316 380 9625
Mailto: Gerd.Leitinger-at-meduni-graz.at




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 04:41:23 2005



From: michael shaffer :      michael-at-shaffer.net
Date: Wed, 9 Mar 2005 07:20:21 -0330
Subject: [Microscopy] RE: RE: RE: SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ken Converse writes ...

} Gary,
} I believe the countrates are correct. If you check
} http://www.rontecusa.com/products.htm
} I think you'll find they have a new detector design.
} It's lN2 free and operates very fast. The resolution is
} not quite as good as their lN2 cooled detectors,
} but the speeds are real.

Our research group chose the Roentec for image analysis applications, but
the data for the 3rd generation Roentec XFlash SDD implies its energy
resolution is better than conventional Si(Li) beginning at input countrates
~80-100kcps. This comparison should be verified against the most recent
Si(Li) electronics announced this last summer. I'll be able to confirm and
provide more details after ours is installed.

cheerios ... shAf :o)
Avalon Peninsula, Newfoundland
www.micro-investigations.com (in progress)



} -----Original Message-----
} } From: Gary Gaugler [mailto:gary-at-gaugler.com]
} Sent: Monday, March 07, 2005 5:55 PM
} To: Evren Cubukcu
} Cc: MSA listserver
} Subject: [Microscopy] RE: SEM - LEO or JEOL?
}
}
}
} ------------------------------------------------------------------
} ----------
} --
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} ------------------------------------------------------------------
} ----------
} ---
}
}
} First of all, what method did you use to arrive
} at the selection of an EVO SEM? Why not high vacuum?
} I think that this needs to be factored if you're not sure.
} What type of final specimen are you going to be analyzing?
} If the soil is embedded, polished low mag imaged, you don't
} necessarily need an ESEM. That said, if this is mostly all you
} will be looking at, you can suffer the loss of resolution
} of an ESEM. That you prefer not needing a chiller
} is important. However, there are many advantages
} in SEM systems that are only achieved by chiller-based
} SEMs. You need to sort out all of the variables
} for a SEM system first, I think, then select an
} appropriate EDS system. In both cases, availability
} of service resources is critical. If you don't have
} a chiller, what cools the diffusion pump or turbo?
} Some turbos can be air cooled.
}
} If you are going to do an ESEM, it may likely
} have a W filament or LaB6 at best. Thus, how
} are you going to achieve 400K cps much less than
} actually processing them? Furthermore, regardless
} of the maximum cps stated, what filter time is this based on?
} Short times can handle large cps at low resolution
} and low fidelity. As filter time increases, dead time
} increases and the existence of many counts becomes
} irrellevant if not counterproductive. Most (my guess)
} systems run about 2000 cps at 30% DT at 102uS time
} constant. If you are concerned about resolution and
} sensitivity, you should focus on obtaining long filter
} time and { 35% DT. A system can certainly handle
} hundreds of thousands cps but not likely at long filter
} times. In this case, short filter times don't produce
} good quant results. However, short times are valuable for
} mapping. So, keep this in mind based on your application.
} If you seek high resolution, you need a very good detector
} and long filter times. Additionally, it seems out of
} character to me that a W system would produce very high
} counts compared to a LaB6 or especially an FESEM. But
} I've only done EDS on FESEMs.
}
} The other issue to keep in mind is that if you want to
} do quant EDS, you will need an EDS package that allows
} for high pressure quant. I use EDAX Genesis VIP Quant
} for variable pressure SEM (usually 5-50Pa). It is selectable
} for VP or high vacuum. Furthermore, it provides for
} ZAF, PhiZAF and PhiRhoZAF quant in VP or HV modes.
} The user interface is very nice, IMO. I have used the
} Rontec as well and got good results in a high vacuum
} system. I don't know if they can handle high pressure
} situations.
}
} There have been discussions about standardless quant
} on this list before. The method I use is what EDAX calls
} SEC--a standards correction factor. This is where the
} quant value is typically value*SEC where SEC=1. However,
} the SEC value can be changed to a value that makes the
} quant result match a standard. Thus, quant results are
} done routinely in a standardless manner but are traced
} back to a standard. This SEC check should be done on
} a routine basis (I do it every quarter). The other thing
} to check is that the peak for each detected element lines
} up with its respective eV for whichever shell is being
} examined. This value can shift slightly. Sometimes,
} an element's eV value might be high or low by a small
} amount and confuse a convolution (peak pileup).
}
} As far as the detector being LN2-free, I agree very much.
} The tradeoffs for this option are cost and performance.
} Without getting at or close to LN2 temperature, detector
} performance will suffer. I use the EDAX Cryospec LN2-free
} detector. This detector produces the same performance
} as an LN2 detector. Currently, I am getting 131.5eV
} resolution and can detect down to and including Be.
} But of course, YMMV.
}
} gary g.
}
} At 12:44 AM 3/7/2005, you wrote:
}
}
} } -----------------------------------------------------------------
} ----------
} ---
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------
} ----------
} ----
} }
} } Thank you very much for such valuable comments and suggestions.
} } I have been in SEM field for 2 years now and your contributions
} are really
} } amazing.
} }
} } Evren
}
}
}
}
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 04:50:53 2005



From: Ken Converse :      kenconverse-at-qualityimages.biz
Date: Wed, 9 Mar 2005 06:00:17 -0500
Subject: [Microscopy] Re: SEM - LEO or JEOL?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Gary,
Rontec also makes an lN2 series of detectors with resolution as good as 129
ev. Different people have different needs. Less resolution isn't
necessarily "junk" and high resolution without properly accounting for a lot
of variables will give you very precise "junk". Besides, if resolution is
so critical, why would you use EDS in the first place? You should be using
WDS on a probe with limited stage motion and light optics.

I have no financial interest. Just service the instruments these things get
bolted to.

Ken Converse

owner 
QUALITY IMAGES
Servicing Scanning Electron Microscopes
Since 1981
16 Creek Rd.
Delta, PA 17314
717-456-5491
Fax 717-456-7996
kenconverse-at-qualityimages.biz
qualityimages.biz


-----Original Message-----
} From: Gary Gaugler [mailto:gary-at-gaugler.com]
Sent: Tuesday, March 08, 2005 10:26 PM
To: Ken Converse







From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 10:50:18 2005



From: Geoff McAuliffe :      mcauliff-at-umdnj.edu
Date: Wed, 09 Mar 2005 11:56:16 -0800
Subject: [Microscopy] Re: viaWWW: Myelin Fixation for EM Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Find a paper, a review article or a book with EM's that you would
like to emulate. Then try their protocol. Without knowing something
about your experimental conditions, protocol and the problems you are
encountering it is difficult to make specific recommendations. Are you
fixing by immersion, perfusion, composition of buffer and fix,
post-fixation, dehydration, etc. etc? Also, keep in mind that optimal
preservation of immuno may not provide optimal perservation for morphology.

Geoff

by way of MicroscopyListserver wrote:

} ------------------------------------------------------------------------------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (mitrah-at-uci.edu) from
} http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
} Tuesday, March 8, 2005 at 17:46:59
} ---------------------------------------------------------------------------
}
}
} Email: mitrah-at-uci.edu
} Name: Mitra Hooshmand
}
} Organization: UC Irvine
}
} Title-Subject: [Microscopy] [Filtered] MListserver: Myelin Fixation
} for EM Analysis
}
} Question: I have been unsuccessfully attempting at obtaining optimal
} myelin morphology for EM analysis. My ultimate goal is to do
} immuno-EM on the spinal cord. However, I have not been able to get
} myelin ultrastructure preservation even in doing EM alone.
}
} If you have any recommendations/suggestions or a protocol that may
} guide me, I would be most grateful. I am on an unforgiving deadline
} and in dire need for some guidance. Thank you, immensely, in advance.
}
} Mitra Hooshmand
} Graduate Student Researcher
} UC Irvine
} mitrah-at-uci.edu
}
} ---------------------------------------------------------------------------
}
}
}

--
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************





From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 13:17:04 2005



From: Jeremy Sanderson :      jb_sanderson-at-yahoo.com
Date: Wed, 9 Mar 2005 19:26:42 +0000 (GMT)
Subject: [Microscopy] Literature on Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello listers,
While following the recent useful thread about costs
of running imaging facilities, I wondered what had
been written about setting up a microscopy imaging
facility.

The most recent reference that I can find is:
DeMaggio, Susan (2002) Running and Setting up a
Confocal Microscope Core Facility. Chapter 14 in:
Methods in Cell Biology 70: 475-485

I've trawled the archives of both listservers, and
maybe I've missed something. Perhaps this stuff isn't
written down. This particular reference seems to have
been written by a successful cytometrist.

Have any of you microscopists out there written
anything, or know of anything, that can be referred to
or cited?

Regards,
Jeremy

Send instant messages to your online friends http://uk.messenger.yahoo.com


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 14:19:03 2005



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Thu, 10 Mar 2005 09:29:08 +1300
Subject: [Microscopy] Point Counter Supplier(s)?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

Does anyone know who, these days, supplies microscope stage point counter
attachments for petrographic use?

We have an old one made by Swift, of England, but my Google searching has not
turned up any current suppliers or manufacturers.

tia

rtch

--
Ritchie Sims Ph D Phone : 64 9 3737599 ext 87713
Microanalyst Fax : 64 9 3737435
Department of Geology email : r.sims-at-auckland.ac.nz
The University of Auckland
Private Bag 92019
Auckland
New Zealand



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 14:51:36 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Wed, 09 Mar 2005 16:00:35 -0500
Subject: [Microscopy] Re: Literature on Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Jeremy,
The only comprehensive source I know of is 30 years old...Its Volume
4 in the "Practical Methods in Electron Microscopy" series edited by
Audrey Glauert: "Design of the Electron Microscope Laboratory" by
Ronald H. Alderson (1975) North-Holland.
I am in the process of planning a relocation of my facility, and this
book is still extremely helpful.
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 15:40:57 2005



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Wed, 09 Mar 2005 16:49:48 -0500
Subject: [Microscopy] Re: Literature on Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Jeremy,
Requirements for imaging facilities vary depending on the type of equipment
it will contain. Also you are often constrained by the location of the
facility.

For example, vibration problems for LM can often be overcome by using
weighted tables. High resolution SEM or TEM often requires very stable
sub-basements or intricate vibration mounts.

One presentation, and related discussion, at the Core Facility Management
session at M&M 2005 will be on setting up an appropriate environment for
high-resolution TEM taking into consideration vibration, electric fields,
temperature shifts, and acoustic problems.

Debby

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy


On 3/9/05 2:26 PM, "Jeremy Sanderson" {jb_sanderson-at-yahoo.com} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Hello listers,
} While following the recent useful thread about costs
} of running imaging facilities, I wondered what had
} been written about setting up a microscopy imaging
} facility.
}
} The most recent reference that I can find is:
} DeMaggio, Susan (2002) Running and Setting up a
} Confocal Microscope Core Facility. Chapter 14 in:
} Methods in Cell Biology 70: 475-485
}
} I've trawled the archives of both listservers, and
} maybe I've missed something. Perhaps this stuff isn't
} written down. This particular reference seems to have
} been written by a successful cytometrist.
}
} Have any of you microscopists out there written
} anything, or know of anything, that can be referred to
} or cited?
}
} Regards,
} Jeremy
}
} Send instant messages to your online friends http://uk.messenger.yahoo.com
}




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 9 19:13:23 2005



From: moss-at-relia.net (by way of MicroscopyListserver)
Date: Wed, 9 Mar 2005 19:23:29 -0600
Subject: [Microscopy] viaWWW: Zeiss 902 manuals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (moss-at-relia.net) from
http://microscopy.com/MicroscopyListserver/MLFormMail.html on
Wednesday, March 9, 2005 at 13:20:41
---------------------------------------------------------------------------

Email: moss-at-relia.net
Name: William McManus

Organization: Mt Ogden Scientific Services

Title-Subject: [Microscopy] [Filtered] Zeiss 902 manuals

Question:
I am working on a Zeiss 902, and need to find a set of tech and
service manuals in english. If anyone has access to any please
contact me.
Bill

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 01:57:31 2005



From: Gareth Morgan :      Gareth.Morgan-at-labmed.ki.se
Date: Thu, 10 Mar 2005 09:07:39 +0100
Subject: [Microscopy] Re: Literature on Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

I would also be interested in other peoples experience in this. I run a
LASER dissection plus an image analysis core facility here and for me it is
just a matter of trial and error (maybe mostly error!?) as a
'sideline/spare time' activity alongside other (significant) teaching
commitments.

See
http://labmed.ki.se/avdelningar/patol/core_facility/corefacility
This is not an advertisement - I cannot imagine that any of you would want
to use us.

All the best

Gareth


With best wishes,

Gareth

http://www.ki.se/biomedlab
e-mail Gareth.Morgan-at-labmed.ki.se

Tel +46 8 5858 1038
Fax +46 8 5858 7730

Gareth Morgan MPhil MSc FIBMS,
Department of Histo/cytopathology, Laboratory Medicine (Labmed),
Karolinska Institute,
Karolinska University Hospital at Huddinge, F46
SE 141 86 Stockholm
Sweden

OBS! Besöksadress: F-Huset, Forskningsgatan 2 F52, Rum 2.10. Laboratoriet
för klinisk patologi och cytologi.

NB! Visiting address: Building F, Research Corridor 2 F52, Room 2.10.
Clinical Histo- and Cytopathology Laboratory.




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 09:38:38 2005



From: Fred Pearson :      eoptics-at-mcmaster.ca
Date: Thu, 10 Mar 2005 10:49:57 -0500 (Eastern Standard Time)
Subject: [Microscopy] Announcement of Microscopical Society of Canada (MSC) Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


To the Microscopy Community:

I would like to draw your attention to the microscopy meeting of the
MSC to be held at McMaster University in Hamilton Ontario Canada in May
2005. The conference will held on May 18-20. Workshops will be on
May 16-17.

We have prepared a superb program covering themes such as:

synchrotron-base microscopy,
electron microscopy (TEM and low voltage/environmental SEM),
tomography of biological and materials structures,
scanning probe techniques,
spectroscopy in the TEM,
focused ion beam microscopy of bio/materials etc.
advances in instrumentation

Prior to the meeting, there will be a series of worskhops jointly held
with the 4th Annual Brockhouse Institute for Materials Research meeting.
The Workshops will include:

cryomicroscopy,
confocal and fluorescence microscopy,
image analysis,
scanning probe methods
EBSD
spectromicroscopy
microtomy
energy loss spectroscopy and energy filtered imaging
high-resolution electron microscopy (quantitative measurement of
strains at defects)
and many more.

Please see
http://www.brockhouse.mcmaster.ca/MSC-SMC2005/
for a full program, speakers, registrations etc.

There will be exhibits from manufacturers in microscopy-related tools
from May 18 to the 20. This is the perfect occasion to hear about
the latest developments and see demonstrations.

There will demos on:
FIB ( with environmental SEM), FE-SEM, EBSD.

The Science Advisor to the Prime Minister of Canada, Art Carty, will be at
the welcoming mixer on Tuesday (joint event with the Brockhouse Institute
Annual workshop mixer).

Please pass this on to your research group, post the announcement in
your laboratories and encourage your group to attend. This is going to
be a really excellent meeting and you should not miss this opportunity
to participate to this event.

Note the Abstract submission deadline 15th March, Registration deadline
15 April. Register early to keep a space in the workshops.
Kind regards to all of you. Hope to see you in May.


Gianluigi Botton, (on behalf of the organising committee)




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 10:28:17 2005



From: Aleksandr Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Thu, 10 Mar 2005 17:37:17 +0000
Subject: [Microscopy] Soda lime glass coverslips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear listers,
I'm forwarding this email that I received in the hope that our vast pool of
knowledge, info and contacts might be put to use again. I would have
started this person out with Peter Stoltzenberg, but as many of you realize,
he is sadly out of the service business. Is there anyone else who is
familiar with the Siemens 1A, preferably in the South? Please respond
directly to Onesimus Otieno at ootieno-at-oaks1.oakwood.edu . Thanks in
advance.

Ken Converse

owner 
QUALITY IMAGES
Servicing Scanning Electron Microscopes
Since 1981
16 Creek Rd.
Delta, PA 17314
717-456-5491
Fax 717-456-7996
kenconverse-at-qualityimages.biz
qualityimages.biz


-----Original Message-----
} From: ootieno-at-oaks1.oakwood.edu [mailto:ootieno-at-oaks1.oakwood.edu]
Sent: Wednesday, March 09, 2005 2:47 PM
To: kenconverse-at-qualityimages.biz

Dear listers,

I desperately need soda lime glass coverslips. I've tried to search the
suppliers on the Interent but without any success. I cannot use the
usual borosilicate coverslips as they seemingly inhibit the gold
enhancement reaction. I cannot use soda lime glass slides because they
are too thick for my purposes.
I will greatly appreciate any advice.

--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 11:36:31 2005



From: K.N. Bozhilov :      bozhilov-at-ucr.edu
Date: Thu, 10 Mar 2005 09:46:26 -0800
Subject: [Microscopy] Defocus spread

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am wondering whether there is alternative to determining the defocus
spread on a HR TEM without an energy filtering capabilities in addition
to the Young's fringe method of Frank and the elaborate measurements of
actual voltage, beam current, and lens current fluctuations.

_____________________________________________
Krassimir N. Bozhilov
Central Facility for Advanced Microscopy and Microanalysis
University of California
Riverside, CA 92521

tel 951 827 2998
fax 951 827 2489
________________________________________________



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 15:07:41 2005



From: Hong Yi :      hyi-at-emory.edu
Date: Thu, 10 Mar 2005 16:17:41 -0500
Subject: [Microscopy] (Microscopy)Re: CCD camera on TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear All:

           We had a Gatan BioScan CCD camera installed on our TEM about
6 years ago when we bought the scope, but for some reason we have not
used the camera much. Now our situation has changed that we will
benefit from using a CCD camera so we want to get it running
again. However, we need to switch our computer from a Mac to a PC and
upgrade the software for the camera. I got a price quotation from the
company already. But my concern is that the model has been around for
so long and technology probably has improved a lot since, so is it
worth it for us to upgrade only software? I would like to hear opinions
from people who have experience with Gatan BioScan and other CCD. We
are a multi-user facility, so the samples we look at on the scope can
be anything from nano magnetic particles to rat brain. 

           Thank you in advance. 

Hong
Emory EM



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 16:47:08 2005



From: Gordon Couger :      gcc-at-couger.com
Date: Thu, 10 Mar 2005 16:56:41 -0600
Subject: [Microscopy] Re: Soda lime glass coverslips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Will plastic cover slips work?

Do your have to equipment to make thin sections for geology
While they would be rather expensive you should be able to make
soda lime cover slips from soda lime slides by grinding them
down and polishing them.

If you have sputtering equipment you might be able to sputter a
coating on regular cover slips that does not inhibit your reaction.

Good luck
Gordon
Gordon Couger

I collect links on information related to light microscopes.
www.couger.com/microscope/links/gclinks.html
Please forward anything you think might be useful to others.
Microscope Documentation is at www.science-info.org

Aleksandr Mironov wrote:
}
}
} ------------------------------------------------------------------------------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
}
} Dear listers,
}
} I desperately need soda lime glass coverslips. I've tried to search the
} suppliers on the Interent but without any success. I cannot use the
} usual borosilicate coverslips as they seemingly inhibit the gold
} enhancement reaction. I cannot use soda lime glass slides because they
} are too thick for my purposes.
} I will greatly appreciate any advice.
}



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 16:51:27 2005



From: Walck, Scott D. :      walck-at-ppg.com
Date: Thu, 10 Mar 2005 18:01:04 -0500
Subject: [Microscopy] Soda lime glass coverslips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I don't think that I am aware of any soda lime coverslips. If you find them, I would be interested in them for purposes of my own.

I don't know what " the gold enhancement reaction" is, but I do have a couple of suggestions.

1) You might consider dipping your borosilicate coverslips slowly into a diluted sodium silicate solution (also known as water glass) and then heating them up afterward ( I would suggest 500 -600C). This might put enough Na into the glass to give you similar properties as what you want. You might even be able to use the water glass directly dipping it onto other substrates or a glass slide.

2) You can use some of the standard mechanical thinning processes to thin float glass to about 100 um thick. Talk to any of the TEM prep guys for the physical sciences, e.g South Bay Technology, Gatan, E. A Fischione, and several others. There are devices that will allow you to make parallel sided samples to either the Sn side or air side of the glass. (In the dark, the Sn side will fluoresce with UVB illumination.) By using standard polishing techniques, you can get your polished side to be very smooth without much difficulty. Talk to some people in the physical sciences there that are doing TEM. I'm sure that they can help you out.

-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472
Walck-at-PPG.com
(412) 820-8651 (office)
(412) 820-8515 (fax)


-----Original Message-----
} From: Aleksandr Mironov [mailto:Aleksandr.Mironov-at-manchester.ac.uk]
Sent: Thursday, March 10, 2005 12:37 PM
To: Microscopy-at-microscopy.com

Dear listers,

I desperately need soda lime glass coverslips. I've tried to search the
suppliers on the Interent but without any success. I cannot use the
usual borosilicate coverslips as they seemingly inhibit the gold
enhancement reaction. I cannot use soda lime glass slides because they
are too thick for my purposes.
I will greatly appreciate any advice.

--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk






From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 17:00:14 2005



From: Dohnalkova, Alice :      Alice.Dohnalkova-at-pnl.gov
Date: Thu, 10 Mar 2005 15:09:34 -0800
Subject: [Microscopy] (Microscopy)Re: CCD camera on TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hi Hong,
We have a Gatan's 1x1K CCD about the same age as yours that is being used heavily (on a daily basis). We are still running the old Mac-based Digital Micrograph for image acquisition, but we do most of the image analyses on a newer PC DM version. Likewise, we image both biological and material science samples, and the camera has proven itself to be a great workhorse piece of equipment.

Best regards,
Alice.


Alice Dohnalkova
Environmental Microbiology
Pacific Northwest National Laboratory
Richland, WA 99352
(509) 372-0692 office
(509) 376-3654 TEM lab

-----Original Message-----
} From: Hong Yi [mailto:hyi-at-emory.edu]
Sent: Thursday, March 10, 2005 1:18 PM
To: Microscopy-at-microscopy.com

Dear All:

           We had a Gatan BioScan CCD camera installed on our TEM about
6 years ago when we bought the scope, but for some reason we have not
used the camera much. Now our situation has changed that we will
benefit from using a CCD camera so we want to get it running
again. However, we need to switch our computer from a Mac to a PC and
upgrade the software for the camera. I got a price quotation from the
company already. But my concern is that the model has been around for
so long and technology probably has improved a lot since, so is it
worth it for us to upgrade only software? I would like to hear opinions
from people who have experience with Gatan BioScan and other CCD. We
are a multi-user facility, so the samples we look at on the scope can
be anything from nano magnetic particles to rat brain. 

           Thank you in advance. 

Hong
Emory EM






From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 18:33:16 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Thu, 10 Mar 2005 16:46:33 -0800
Subject: [Microscopy] Re: (Microscopy)Re: CCD camera on TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Hong
We have Gatan's 600W BioScan camera for about 3 years and very happy. This
is very good camera for most applications. Because it's top-mount camera,
you could not expect "atomic" resolution on this camera. From another hand
you have bigger area of view, which sometime is very useful especially in
biological applications. For really fine works (material science mostly)
you need bottom-mount camera like UltraScan. Basically, (if you have
enough money and want "go digital") you need to have both: bottom and top
mounted cameras. Both would be operated from the same software
(DigitalMicrograph), which is quite convenient. I do find
DigitalMicrograph (DM) software quite good and easy to manipulate. I don't
like the way, how you export images into the TIFF with DM - a lot of mouse
clicks and I was not able to create workable macros to automate the job
(16-bit images). You also lost scale bar... Finally I gave up and export
only for publications (which I believe was intention of DM
creators). Anyway, I am quite happy user and have no interest in Gatan
unless I'll find money for UltraScan.
Sergey

At 04:17 PM 3/10/2005 -0500, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 20:01:55 2005



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Fri, 11 Mar 2005 15:12:05 +1300
Subject: [Microscopy] Point Counters

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

In case anyone else needs to know, I am told that Swift doesn't make point counters
any longer, but that Leica does.


cheers

rtch

--
Ritchie Sims Ph D Phone : 64 9 3737599 ext 87713
Microanalyst Fax : 64 9 3737435
Department of Geology email : r.sims-at-auckland.ac.nz
The University of Auckland
Private Bag 92019
Auckland
New Zealand



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 10 21:58:28 2005



From: Gazda, Jerzy :      jerzy.gazda-at-ceriumlabs.com
Date: Thu, 10 Mar 2005 22:07:49 -0600
Subject: [Microscopy] (Microscopy)Re: CCD camera on TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hong,
We have two microscopes equipped with Gatan MSC794 CCD cameras, one of them had a PC interface for the last six years, the second one came originally on MAC running DM2.5. Two years ago we switched to IEEE1394 Firewire control and modern PC operations on both of these cameras. The hardware cost and software cost was in my opinion minimal as compared to purchases of brand new cameras and that was not even considered.

Having said that, going form Mac to PC had required learning new habits and dealing with few bugs in scripts and DM software. Overall the transition was painless and benefits of having modern computers and an upgrade path for both hardware and software, priceless. We do not use film routinely and on one of the microscopes never had.

Regards,

Jerzy
******************************************************
Jerzy Gazda, Ph.D. CeriumLaboratories L.L.C.
Supervising Engineer 5204 E. Ben White Blvd. - MS 512
Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453
jerzy.gazda-at-ceriumlabs.com
******************************************************




-----Original Message-----
} From: Hong Yi [mailto:hyi-at-emory.edu]
Sent: Thursday, March 10, 2005 3:18 PM
To: Microscopy-at-microscopy.com

Dear All:

           We had a Gatan BioScan CCD camera installed on our TEM about
6 years ago when we bought the scope, but for some reason we have not used the camera much. Now our situation has changed that we will benefit from using a CCD camera so we want to get it running again. However, we need to switch our computer from a Mac to a PC and upgrade the software for the camera. I got a price quotation from the company already. But my concern is that the model has been around for so long and technology probably has improved a lot since, so is it worth it for us to upgrade only software? I would like to hear opinions from people who have experience with Gatan BioScan and other CCD. We are a multi-user facility, so the samples we look at on the scope can be anything from nano magnetic particles to rat brain. 

           Thank you in advance. 

Hong
Emory EM







From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 00:46:15 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Thu, 10 Mar 2005 22:55:36 -0800
Subject: [Microscopy] Re: (Microscopy)Re: CCD camera on TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sergey;
Gatan has a plug-in that will convert images to any format you
want. Two mouse clicks will convert as many images as you can fit in a
folder. Sound good? You will love it. Also, DM 3.9 has the converter
built-in.

John Mardinly
Intel

-----Original Message-----
} From: Sergey Ryazantsev [mailto:sryazant-at-ucla.edu]
Sent: Thursday, March 10, 2005 4:47 PM
To: Microscopy-at-microscopy.com

Dear Hong
We have Gatan's 600W BioScan camera for about 3 years and very happy.
This
is very good camera for most applications. Because it's top-mount
camera,
you could not expect "atomic" resolution on this camera. From another
hand
you have bigger area of view, which sometime is very useful especially
in
biological applications. For really fine works (material science mostly)

you need bottom-mount camera like UltraScan. Basically, (if you have
enough money and want "go digital") you need to have both: bottom and
top
mounted cameras. Both would be operated from the same software
(DigitalMicrograph), which is quite convenient. I do find
DigitalMicrograph (DM) software quite good and easy to manipulate. I
don't
like the way, how you export images into the TIFF with DM - a lot of
mouse
clicks and I was not able to create workable macros to automate the job
(16-bit images). You also lost scale bar... Finally I gave up and export

only for publications (which I believe was intention of DM
creators). Anyway, I am quite happy user and have no interest in Gatan
unless I'll find money for UltraScan.
Sergey

At 04:17 PM 3/10/2005 -0500, you wrote:


} -----------------------------------------------------------------------
-------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America

} the camera. I got a price quotation from the company already. But my
} concern is that the model has been around for so long and technology
} probably has improved a lot since, so is it worth it for us to upgrade

} only software? I would like to hear opinions from people who have
} experience with Gatan BioScan and other CCD. We are a multi-user
} facility, so the samples we look at on the scope can be anything from
} nano magnetic particles to rat brain.
}
} Thank you in advance.
}
} Hong
} Emory EM
}





From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 01:25:43 2005



From: akoorts-at-medic.up.ac.za (by way of Ask-A-Microscopist)
Date: Fri, 11 Mar 2005 01:35:51 -0600
Subject: [Microscopy] AskAMicroscopist: Prussian Blue iron stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (akoorts-at-medic.up.ac.za) from
http://www.msa.microscopy.com/Ask-A-Microscopist/Ask-A-Microscopist.html
on Friday, March 11, 2005 at 00:03:53
---------------------------------------------------------------------------

Email: akoorts-at-medic.up.ac.za
Name: Alida Koorts

Organization: University of Pretoria

Education: Graduate College

Location: Pretoria, South Africa

Question: Is it possible to perform a Prussian Blue iron stain on LR
White sections for light microscopy. If so I would very much
appreciate a detailed protocol.
Best regards
Alida

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 07:16:12 2005



From: michael shaffer :      michael-at-shaffer.net
Date: Fri, 11 Mar 2005 09:56:19 -0330
Subject: [Microscopy] printing round labels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I hope this doesn't deviate too far from the scope of this list ... but I am
having problems finding what appears to be a unique label printer. As part
of a high-throughput sample preparation workflow, we need to print
individual 1" round labels. What methods are others using?

cheerios ... shAf :o)
Avalon Peninsula, Newfoundland
{www.micro-investigations.com}



From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 09:10:47 2005



From: Ellery Frahm :      frah0010-at-tc.umn.edu
Date: Fri, 11 Mar 2005 09:49:58 -0600
Subject: [Microscopy] magnetic lenses; classroom example

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Michael
Avery produce round labels for laser and inkjet printing, and the software
for printing them using various wordprocessors
using standard printers
see e.g.

http://www.avery.com/us/Main?action=product.HierarchyList&node=10211270&catalogcode=WEB01

Chris

----- Original Message -----
} From: "michael shaffer" {michael-at-shaffer.net}
To: "Microscopy list" {Microscopy-at-MSA.Microscopy.Com}
Sent: Friday, March 11, 2005 1:26 PM

Microscopy folks,

I am always looking for interesting visual aids for the course I teach
on scanning electron microscopy and X-ray microanalysis. For instance,
I have an old WD spectrometer that I bring to lecture, let the student
turn the screw that moves the dispersing crystal and detector, etc.

For lectures on magnetic lenses, I'd like to use either unwound
windings from an old magnetic lens or just an equivalent length and
gauge of copper wire. Unfortunately, it is difficult to find online
suppliers of old or dead electron microscopy components (not much
demand, I suspect), and I cannot find any references (or even rough
estimates) to the number of windings in, say, a typical condenser lens
or what the gauge of that wire would be.

Any help with either of these endeavors would be much appreciated.

Thanks,
Ellery

--------------------
Ellery E. Frahm
Research Fellow & Manager
Electron Microprobe Laboratory
University of Minnesota - Twin Cities
Department of Geology & Geophysics
Lab Website: http://probelab.geo.umn.edu



From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 10:56:41 2005



From: Michael O'Keefe :      MAOK-at-lbl.gov
Date: Fri, 11 Mar 2005 09:06:26 -0800
Subject: [Microscopy] [Fwd: Defocus spread]

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Michael O'Keefe wrote:

} Krassimir,
}
} You mention two methods of determining defocus spread. Like you, I'd also be
} interested to know of any third way.
}
} 1. Defocus spread is a function of beam energy spread plus variations
} (instabilities) in lens current and high voltage times the coefficient of chromatic
} aberration -- so you can measure all these and compute the defocus spread.
} Problems include getting access to the lens current and measuring fluctuations of
} the order of one part per million, measuring beam energy spread (with a filter you
} can measure the energy spread from the tip with the combined HT instabilities by
} integrating the zero-loss peak over a typical image collection time -- this is
} safer than trying to get access to the HT (preferably at the gun) and measuring
} parts per million at several hundred kV).
}
} 2. Microscope information limit (where the Young's fringes die out) depends on
} defocus spread -- so you can get an experimental measurement (estimate?) of the
} defocus spread by measuring the extent of the fringes. Problems include the
} specimen (very thin, amorphous, good scatterer) and exact magnification
} calibration. Also, Young's fringes from a thicker specimen can theoretically
} extend beyond the information limit.
}
} Mike
}
} ------------------------
} Dr. Michael A. O’Keefe
} Director, Microscopy Society of America
} Lawrence Berkeley National Laboratory
} Materials Sciences Division
} 1 Cyclotron Road
} Berkeley, CA 94720
} 510-486-4610
} 510-486-5530 fax
} maok-at-lbl.gov
}
} "K.N. Bozhilov" wrote:
}
} } ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } -------------------------------------------------------------------------------
} }
} } I am wondering whether there is alternative to determining the defocus
} } spread on a HR TEM without an energy filtering capabilities in addition
} } to the Young's fringe method of Frank and the elaborate measurements of
} } actual voltage, beam current, and lens current fluctuations.
} }
} } _____________________________________________
} } Krassimir N. Bozhilov
} } Central Facility for Advanced Microscopy and Microanalysis
} } University of California
} } Riverside, CA 92521
} }
} } tel 951 827 2998
} } fax 951 827 2489
} } ________________________________________________



From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 12:43:25 2005



From: pgrover :      pgrover-at-bilbo.bio.purdue.edu
Date: Fri, 11 Mar 2005 13:53:38 -0500
Subject: [Microscopy] re: magnetic lenses; classroom example

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The EM center in agriculture at Purdue had a very nicely made bisected
column showing the condensor lens windings in cross section, truly inspiring
to look at. I remember counting wires in a given area under a stereoscope
and calculating the length of the coil. Debby Sherman probably could tell
you more. I have a couple of old columns I've salvaged and would someday
like to do likewise, but it probably would not be easy, requiring a good
metal cutting bandsaw, then potting the exposed lens in epoxy, then surface
polishing etc. But I guess it'd be about as easy to make two or four
hemi-columns as it would be to make one. I'd be happy to send one of them
to a good home.

Paul Grover
-------------------------------------------------------------------------
"'Why listen, Lady,' he said with a grin of delight 'the monks of old
slept in their coffins!' 'They wasn't as advanced as we are.' the old
woman said." - The Life You Save May be Your Own


----------------------------------------------------------------------------
--
The Microscopy ListServer -- Sponsor: The Microscopy Society of America To
Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver

Microscopy folks,

I am always looking for interesting visual aids for the course I teach
on scanning electron microscopy and X-ray microanalysis. For instance,
I have an old WD spectrometer that I bring to lecture, let the student
turn the screw that moves the dispersing crystal and detector, etc.

For lectures on magnetic lenses, I'd like to use either unwound
windings from an old magnetic lens or just an equivalent length and
gauge of copper wire. Unfortunately, it is difficult to find online
suppliers of old or dead electron microscopy components (not much
demand, I suspect), and I cannot find any references (or even rough
estimates) to the number of windings in, say, a typical condenser lens
or what the gauge of that wire would be.

Any help with either of these endeavors would be much appreciated.

Thanks,
Ellery

--------------------
Ellery E. Frahm
Research Fellow & Manager
Electron Microprobe Laboratory
University of Minnesota - Twin Cities
Department of Geology & Geophysics
Lab Website: http://probelab.geo.umn.edu





From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 12:51:04 2005



From: Quinn, Tim L. :      quinntl-at-umkc.edu
Date: Fri, 11 Mar 2005 12:57:39 -0600
Subject: [Microscopy] mouse serum blocking solution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi gang!

I need to make up a batch of mouse serum blocking buffer to block free biotinylation used on biotintylated mouse Ab on rat tissue.

Does anyone have a recipe mouse serum blocking buffer?

Cheers,

Timothy L. Quinn
Senior Lab Technician
UMKC Medical School
Med Lab
Laboratory of Pulmonary &
Infectious Disease
M3-202
2411 Holmes St
Kansas City, MO 64108-2792

Phone: 816-235-1904
Fax: 816-235-5194
E-mail: quinntl-at-umkc.edu



From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 14:28:14 2005



From: Geoff McAuliffe :      mcauliff-at-umdnj.edu
Date: Fri, 11 Mar 2005 15:40:32 -0800
Subject: [Microscopy] Re: AskAMicroscopist: Prussian Blue iron stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Greetings Alida:

Since immunostaining reagents can penetrate LR White, I would think that
the reagents for Prussian blue would as well. This is a very simple
"stain" (a reaction really), any histotechnique text or website will
have it. It is VERY important the the HCl-potassium ferrocyanide
solution be mixed immediately before use.

Geoff

by way of Ask-A-Microscopist wrote:

} ------------------------------------------------------------------------------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was
} submitted by (akoorts-at-medic.up.ac.za) from
} http://www.msa.microscopy.com/Ask-A-Microscopist/Ask-A-Microscopist.html
} on Friday, March 11, 2005 at 00:03:53
} ---------------------------------------------------------------------------
}
}
} Email: akoorts-at-medic.up.ac.za
} Name: Alida Koorts
}
} Organization: University of Pretoria
}
} Education: Graduate College
}
} Location: Pretoria, South Africa
}
} Question: Is it possible to perform a Prussian Blue iron stain on LR
} White sections for light microscopy. If so I would very much
} appreciate a detailed protocol.
} Best regards
} Alida
}
} ---------------------------------------------------------------------------
}
}
}

--
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************





From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 18:35:52 2005



From: Jan Leunissen :      leunissen-at-aurion.nl
Date: Sat, 12 Mar 2005 13:45:15 +1200
Subject: [Microscopy] Immunoreagents penetrating into LR White?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Kia Ora (albeit with a Dutch accent),

Has anyone observed that immunoreagents actually penetrate into
LR-white sections? It seems so unlikely to me. Antibodies being 8-12nm
in size, not even taking into account the size increase caused by any
marker attached....... Wouldn't an ultrathin section, that is anything
between let's say 40 and 70 nm thick, look like a slice of Gouda Cheese
instead of a smooth and even layer? Depending on fixation and maybe
other treatments penetration into even fully hydrated ultrathin
cryosections is usually limited to antibodies and the very small gold
particle immunoconjugates, larger ones not getting deeper than the
surface. York-Dieter Stierhof from Tübingen, Germany, did some
beautiful work on this topic.
I am really interested in this and would appreciate to hear of anyone's
experience.

BTW, I have no commercial interest in Gouda Cheese...

Jan Leunissen
Aurion Present Address:
Costerweg 5 EM-Unit
6702 A Wageningen Otago School of Medicine
The Netherlands Dunedin, New Zealand
phone 31-317-497676 phone 64-3-4797109
fax 31-317-415955 fax 64-3-4797254
http://www.aurion.nl http://ocem.otago.ac.nz
------------------------------------------------------------------------
--------
"Light Microscopy? That is for people on a diet ". . . Eva Leunissen
(12)


On Mar 12, 2005, at 11:40 am, Geoff McAuliffe wrote:

} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Greetings Alida:
}
} Since immunostaining reagents can penetrate LR White, I would think
} that the reagents for Prussian blue would as well. This is a very
} simple "stain" (a reaction really), any histotechnique text or website
} will have it. It is VERY important the the HCl-potassium ferrocyanide
} solution be mixed immediately before use.
}
} Geoff
}
} by way of Ask-A-Microscopist wrote:
}
} } ----------------------------------------------------------------------
} } --------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } America
} } To Subscribe/Unsubscribe --
} } http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } ----------------------------------------------------------------------
} } ---------
} } Below is the result of your feedback form (NJZFM-ultra-55). It was
} } submitted by (akoorts-at-medic.up.ac.za) from
} } http://www.msa.microscopy.com/Ask-A-Microscopist/Ask-A-
} } Microscopist.html on Friday, March 11, 2005 at 00:03:53
} } ----------------------------------------------------------------------
} } -----
} }
} } Email: akoorts-at-medic.up.ac.za
} } Name: Alida Koorts
} }
} } Organization: University of Pretoria
} }
} } Education: Graduate College
} }
} } Location: Pretoria, South Africa
} }
} } Question: Is it possible to perform a Prussian Blue iron stain on LR
} } White sections for light microscopy. If so I would very much
} } appreciate a detailed protocol.
} } Best regards
} } Alida
} }
} } ----------------------------------------------------------------------
} } -----
} }
} }
}
} --
} --
} **********************************************
} Geoff McAuliffe, Ph.D.
} Neuroscience and Cell Biology
} Robert Wood Johnson Medical School
} 675 Hoes Lane, Piscataway, NJ 08854
} voice: (732)-235-4583; fax: -4029 mcauliff-at-umdnj.edu
} **********************************************
}
}
}
}
}
}




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 22:51:32 2005



From: kellymariegreen-at-hotmail.com (by way of Ask-A-Microscopist)
Date: Fri, 11 Mar 2005 23:01:43 -0600
Subject: [Microscopy] AskAMicroscopist: lectures on electron microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (kellymariegreen-at-hotmail.com) from
http://www.msa.microscopy.com/Ask-A-Microscopist/Ask-A-Microscopist.html
on Friday, March 11, 2005 at 08:27:35
---------------------------------------------------------------------------

Email: kellymariegreen-at-hotmail.com
Name: Kelly Green

Organization: UWE

Education: Graduate College

Location: Bristol, UK

Question: After several lectures on electron microscopy (including
sem, tem and edx) i am still very confused. Are there any good
reviews to start me off with the methods and biomedical diagnostic
applications?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 11 22:53:13 2005



From: echeung-at-eyetech.com (by way of MicroscopyListserver)
Date: Fri, 11 Mar 2005 23:03:03 -0600
Subject: [Microscopy] viaWWW: removing glass coverslips from cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (echeung-at-eyetech.com) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
Friday, March 11, 2005 at 17:29:24
---------------------------------------------------------------------------

Email: echeung-at-eyetech.com
Name: Eunice Cheung

Organization: eyetech

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi.
I was wondering if there were any available step by step protocols on
removing glass coverslips from cells grown on them. I have been
trying to study the en face structure of cells. 1) growing them on
coverslips 2)Processing them with Spurrs for ultramicronomy 3) using
BEEM capsules filled with Spurrs resin and inverting them in cotact
with the glass coverslips 4) After polymerizing, using heat to melt
and remove the coverslip. 5) Section and view them on TEM. So far,
the surface of the cells have been stuck to the coverslip after
removal or glass pieces stuck to the sample.
I am open to suggestions for any other strategy.
Thanks.
Eunice


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sat Mar 12 08:30:23 2005



From: Tom Phillips :      phillipst-at-missouri.edu
Date: Sat, 12 Mar 2005 10:44:00 -0600
Subject: [Microscopy] Fwd: viaWWW: removing glass coverslips from cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

You might try hydrofluoric acid. I have used it to remove coverslips
from cell cultures embedded in epoxy resin. As always, use extreme
caution when using acids/bases.
Randy

-----Original Message-----
} From: by way of MicroscopyListserver [mailto:echeung-at-eyetech.com]
Sent: Friday, March 11, 2005 11:03 PM
To: microscopy-at-ns.microscopy.com

I do this all the time with an epoxy (EmBed812) resin. after a short
polymerization (about 7-8 hrs) at 60 C, I slowly immerse the coverslip into
liquid nitrogen and the resin pops off. if i polymerize longer, the bond
between glass and plastic is sometimes too strong and the glass fractures
such that it sticks with the block and prevents sectioning. After
detaching the coverslip, I return the block to the oven for further
polymerization. I recommend you stay away from HF; it is nasty stuff and
should not be needed in this application. good luck.



} ----------
}
} Email: echeung-at-eyetech.com
} Name: Eunice Cheung
}
} Organization: eyetech
}
} Title-Subject: [Microscopy] [Filtered] MListserver:
}
} Question: Hi.
} I was wondering if there were any available step by step protocols on
} removing glass coverslips from cells grown on them. I have been trying to
} study the en face structure of cells. 1) growing them on coverslips
} 2)Processing them with Spurrs for ultramicronomy 3) using BEEM capsules
} filled with Spurrs resin and inverting them in cotact with the glass
} coverslips 4) After polymerizing, using heat to melt and remove the
} coverslip. 5) Section and view them on TEM. So far, the surface of the
} cells have been stuck to the coverslip after removal or glass pieces stuck
} to the sample.
} I am open to suggestions for any other strategy.
} Thanks.
} Eunice
}
}
} ---------------------------------------------------------------------------

Thomas E. Phillips, PhD
Professor of Biological Sciences
Director, Molecular Cytology Core
3 Tucker Hall
University of Missouri
Columbia, MO 65211-7400

573-882-4712 (office)
573-882-0123 (fax)
PhillipsT-at-missouri.edu




From MicroscopyL-request-at-ns.microscopy.com Sat Mar 12 19:32:23 2005



From: Judy Murphy :      murphyjudy-at-comcast.net
Date: Sat, 12 Mar 2005 17:41:32 -0800
Subject: [Microscopy] Re: Literature on Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

If you are looking for articles on designing/setting up microscopy
facilities, I wrote one a couple of years ago for Microscopy Today.
Previous to that one of my articles also appears in Protocols for
Electron Microscopy by John Wiley.

If you cannot find it in the archives for Microscopy Today, let me know.

Murphy, Judy A., 2002. Designing A Microscopy/Analytical
Instrumentation Facility: Step By Step Procedure, Microscopy Today,
Nov., 2002.

Murphy, Judy A., 1993. Designing A Microscopy Facility: Step by Step
Procedure. Procedures in Electron Microscopy 7:5.1-7:5.31, John Wiley
and Sons, New York.


along the same thread
Murphy, Judy A., 2001. Image Management For A Multi-Instrument,
Multi-Platform Microscopy Facility, Scanning, May,2001.

There are several others I wrote, but these should get you started.

Cheers,
Judy



Judy Murphy, PhD
Microscopy, Imaging & Lab Design Consultant
Stockton, CA 95219
murphyjudy-at-comcast.net







Jeremy Sanderson wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


From MicroscopyL-request-at-ns.microscopy.com Sun Mar 13 10:54:16 2005



From: john hoffpauir :      hoffpajo-at-yahoo.com
Date: Sun, 13 Mar 2005 09:03:38 -0800 (PST)
Subject: [Microscopy] Re: viaWWW: removing glass coverslips from cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

well you pretty much have it figured out. the only
suggestion i have is to use araldite(sp?). it is
slightly softer and the coverslips are more easily
removed.
good luck
john
--- by way of MicroscopyListserver
{echeung-at-eyetech.com} wrote:
}
}
}
------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The
} Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help
}
http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
-------------------------------------------------------------------------------
}
} Below is the result of your feedback form
} (NJZFM-ultra-55). It was
} submitted by (echeung-at-eyetech.com) from
}
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html
} on
} Friday, March 11, 2005 at 17:29:24
}
---------------------------------------------------------------------------
}
} Email: echeung-at-eyetech.com
} Name: Eunice Cheung
}
} Organization: eyetech
}
} Title-Subject: [Microscopy] [Filtered] MListserver:
}
} Question: Hi.
} I was wondering if there were any available step by
} step protocols on
} removing glass coverslips from cells grown on them.
} I have been
} trying to study the en face structure of cells. 1)
} growing them on
} coverslips 2)Processing them with Spurrs for
} ultramicronomy 3) using
} BEEM capsules filled with Spurrs resin and inverting
} them in cotact
} with the glass coverslips 4) After polymerizing,
} using heat to melt
} and remove the coverslip. 5) Section and view them
} on TEM. So far,
} the surface of the cells have been stuck to the
} coverslip after
} removal or glass pieces stuck to the sample.
} I am open to suggestions for any other strategy.
} Thanks.
} Eunice
}
}
}
---------------------------------------------------------------------------
}
}



__________________________________
Do you Yahoo!?
Yahoo! Small Business - Try our new resources site!
http://smallbusiness.yahoo.com/resources/


From MicroscopyL-request-at-ns.microscopy.com Sun Mar 13 14:26:06 2005



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Mon, 14 Mar 2005 09:36:15 +1300
Subject: [Microscopy] more on point counters

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

in the interests of completeness, and in case there are others interested........

My attention has been drawn also to www.petrog.com, a site which advertises not only a
point-counting stage, but a digital image-analysing suite for optical-microscopical
petrography.

While I appreciate the replies I have had on this subject off-list, I do wish that replies to
postings in general would be made on-list, so that everyone can read the answers as
well as the questions.

I know that advertising is banned, but when a lister asks if a such-and-such is available,
surely it is reasonable enough, and useful, for a supplier to reply ON-LIST to say that
the such-and-such is available from them, isn't it?

cheers

rtch

--
Ritchie Sims Ph D Phone : 64 9 3737599 ext 87713
Microanalyst Fax : 64 9 3737435
Department of Geology email : r.sims-at-auckland.ac.nz
The University of Auckland
Private Bag 92019
Auckland
New Zealand



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 08:40:53 2005



From: Sherwood, Margaret :      MSHERWOOD-at-PARTNERS.ORG
Date: Mon, 14 Mar 2005 09:51:22 -0500
Subject: [Microscopy] viaWWW: removing glass coverslips from cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have always used liquid nitrogen to remove the cover glass. I drop the whole
capsule (with cover glass) into liquid nitrgon, wait a few seconds, remove (with
tongs) and the cover glass ususally just pops off, the cells remain on the
capsule.



Peggy Sherwood
Lab Associate, Photopathology
Wellman Center for Photomedicine (W224)
Massachusetts General Hospital
55 Fruit Street
Boston, MA 02114
617-724-4839 (voice mail)
617-726-6983 (lab)
617-726-3192 (fax)
msherwood-at-partners.org



-----Original Message-----
} From: echeung-at-eyetech.com [mailto:echeung-at-eyetech.com]
Sent: Saturday, March 12, 2005 12:03 AM
To: microscopy-at-ns.microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (echeung-at-eyetech.com) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
Friday, March 11, 2005 at 17:29:24
---------------------------------------------------------------------------

Email: echeung-at-eyetech.com
Name: Eunice Cheung

Organization: eyetech

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi.
I was wondering if there were any available step by step protocols on
removing glass coverslips from cells grown on them. I have been
trying to study the en face structure of cells. 1) growing them on
coverslips 2)Processing them with Spurrs for ultramicronomy 3) using
BEEM capsules filled with Spurrs resin and inverting them in cotact
with the glass coverslips 4) After polymerizing, using heat to melt
and remove the coverslip. 5) Section and view them on TEM. So far,
the surface of the cells have been stuck to the coverslip after
removal or glass pieces stuck to the sample.
I am open to suggestions for any other strategy.
Thanks.
Eunice


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 08:47:42 2005



From: Ken Converse :      kenconverse-at-qualityimages.biz
Date: Mon, 14 Mar 2005 09:57:15 -0500
Subject: [Microscopy] magnetic lenses; classroom example

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ellery,
Probably the reason you can't find info on the windings is because the
critical term (after the lens design itself) is ampere-turns. If you look
at some of the very old EMs that used vacuum tubes, the lens supplies were
likely to operate at voltages in excess of 100VDC but currents in the mA
range. They used thousands of turns of very fine wire. Solid state EMs
tend to operate at much lower voltages (24-48VDC) and considerably higher
currents (several amps) so they have far fewer turns of heavier gauge wire.
In other words 10,000 turns at 100 mA will generate roughly the same
magnetic field as 300 turns at 3.3 A. In each case you have 1000
ampere-turns. The wire gauge is determined primarily by how small a
diameter you can get away with given the heat generation at the given
current and what your space constraints are. Some lenses are water or oil
cooled so that finer wire can be used due to space limitations.

It sounds like Paul Grover is offering a complete column, which I can't do
at the present time. If the column falls through, I might be able to offer
an objective or condenser lens.

Some years ago I read someone's story on cross-sectioning an EM column, but
I don't recall where I read it. The main point was that they had completely
embedded the column in epoxy (or other clear liquid material), THEN
sectioned and polished. It's a major project requiring a serious powerfeed
bandsaw and a lot of polishing, but the results would be stunning.

Good luck and post some images if you do it!

Ken Converse

owner 
QUALITY IMAGES
Servicing Scanning Electron Microscopes
Since 1981
16 Creek Rd.
Delta, PA 17314
717-456-5491
Fax 717-456-7996
kenconverse-at-qualityimages.biz
qualityimages.biz


-----Original Message-----
} From: Ellery Frahm [mailto:frah0010-at-tc.umn.edu]
Sent: Friday, March 11, 2005 10:50 AM
To: message to:
Cc: Ellery Frahm

Microscopy folks,

I am always looking for interesting visual aids for the course I teach
on scanning electron microscopy and X-ray microanalysis. For instance,
I have an old WD spectrometer that I bring to lecture, let the student
turn the screw that moves the dispersing crystal and detector, etc.

For lectures on magnetic lenses, I'd like to use either unwound
windings from an old magnetic lens or just an equivalent length and
gauge of copper wire. Unfortunately, it is difficult to find online
suppliers of old or dead electron microscopy components (not much
demand, I suspect), and I cannot find any references (or even rough
estimates) to the number of windings in, say, a typical condenser lens
or what the gauge of that wire would be.

Any help with either of these endeavors would be much appreciated.

Thanks,
Ellery

--------------------
Ellery E. Frahm
Research Fellow & Manager
Electron Microprobe Laboratory
University of Minnesota - Twin Cities
Department of Geology & Geophysics
Lab Website: http://probelab.geo.umn.edu







___________________________________________________________
$0 Web Hosting with up to 120MB web space, 1000 MB Transfer
10 Personalized POP and Web E-mail Accounts, and much more.
Signup at www.doteasy.com




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 09:26:43 2005



From: Hong Yi :      hyi-at-emory.edu
Date: Mon, 14 Mar 2005 10:36:37 -0500
Subject: [Microscopy] (Microscopy) Victawet grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thank you, everyone, for replying to my posting on CCD camera for TEM.
I now have a totally unrelated question.

I got several replies to my previous posting on carbon coated grids.
One of them was about treating the grids with Victawet. I would like
to try this method but the cost is somewhat high if we order from a
vender. So I would like to know more before spending money. I would
very much appreciate it if I can hear the experience from those who
have used this type of grids. Thank you.

HOng
Emory EM



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 09:42:58 2005



From: willis.robert-at-epamail.epa.gov
Date: Mon, 14 Mar 2005 10:51:39 -0500
Subject: [Microscopy] SEM: fiber analysis question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am analyzing dust samples for a specific fiber which has been
suggested as a marker for a particular source of contamination.
Ultimately, I want to relate the concentration of the marker fiber to
the amount of contamination of the dust sample (in weight percent). The
marker fibers vary enormously in length from { 10 microns to } 600
microns, and thus span a huge range in mass. My question is: what
measure of fiber concentration would make the most sense in trying to
correlate fiber concentration with the level of contamination? E.g.,
number concentration (per gram of dust), total fiber length per gram of
dust, total area, or total fiber mass per gram of dust? Fiber
concentration on a mass basis is enormously skewed by a small number of
very long fibers, so I'm thinking that number concentration or total
fiber length may be more appropriate.

Thank you for your suggestions.

******************************************************
Robert Willis, Ph.D.
Science Advisor
Alion Science and Technology
c/o U.S. EPA
Mail Drop E-205-06
Research Triangle Park, NC 27711
Tel: 919-541-2809 Fax: 919-541-3566
willis.robert-at-epa.gov
******************************************************




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 11:09:02 2005



From: Battjes, Kevin :      battjes-at-impactanalytical.com
Date: Mon, 14 Mar 2005 12:18:40 -0500
Subject: [Microscopy] MMM Spring Meeting - Call for Papers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Announcement and Call for Papers
Michigan Microscopy &Microanalysis Society Spring Meeting in 2005

Comfort Inn/University Park Conference Center
Mt. Pleasant, MI
May 20th, 2005

Abstract Deadline: April 22nd, 2005

The Spring Meeting of the Michigan Microscopy and Microanalysis Society will
be held on May 20th, 2005 at the Comfort Inn/University Park Conference
Center in Mt. Pleasant, Michigan. In this one-day conference, there will be
two sessions. One is a platform session having approximately 8-10 speakers
representing industry, academia, and research laboratories. The other is a
poster session. In addition to the speakers and posters, vendors will
exhibit a wide range of products and services of interest to the microscopy
community. Presentations are being solicited from researchers in the
Physical and Biological Sciences, including one vendor presentation and an
invited speaker. Student participation is particularly encouraged. Also,
vendors are encouraged to contact the below address to reserve space for
product display.

Abstract Submission
Please submit a 300 to 350 word abstract by April 22nd indicating which
session you prefer (poster or presentation) to:

Geoff Williams
217 Brooks Hall
Biology Department
Central Michigan University
Mt. Pleasant, MI 48859
Ph 989 774 3576
Fax 989 774 3462
Email: ge.willi-at-cmich.edu {mailto:ge.willi-at-cmich.edu}




Kevin Battjes
Impact Analytical Voice 989-832-5555, ext 556
Michigan Molecular Institute Fax 989-832-5560
1910 W. St Andrews Road e-mail: battjes-at-mmi.org
Midland MI 48640 battjes-at-impactanalytical.com




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 12:28:42 2005



From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Mon, 14 Mar 2005 12:37:50 -0600
Subject: [Microscopy] viaWWW: removing glass coverslips from cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Eunice,

Another alternative, if someone hasn't already mentioned it, is to grow
the cells on Thermonox plastic coverslips and embed the coverslips
directly into your resin. You can section these like any other sample
and avoid having to remove a glass cover slip from resin. Thermonox
cover slips are designed for cell culture, should be readily available
the various scientific and EM supply houses and come in a variety of
sizes and shapes.

The drawback is that you will most likely have to section the
coverslips, and thus the cells, in cross-section, rather than having the
nice top-down (or bottoms-up, rather) view of the cell. If that's not a
problem, Thermonox is a great help.

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility---We Do Small Well!
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.emc.missouri.edu









-----Original Message-----
} From: by way of MicroscopyListserver [mailto:echeung-at-eyetech.com]
Sent: Friday, March 11, 2005 11:03 PM
To: microscopy-at-ns.microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (echeung-at-eyetech.com) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
Friday, March 11, 2005 at 17:29:24
------------------------------------------------------------------------
---

Email: echeung-at-eyetech.com
Name: Eunice Cheung

Organization: eyetech

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi.
I was wondering if there were any available step by step protocols on
removing glass coverslips from cells grown on them. I have been trying
to study the en face structure of cells. 1) growing them on coverslips
2)Processing them with Spurrs for ultramicronomy 3) using BEEM capsules
filled with Spurrs resin and inverting them in cotact with the glass
coverslips 4) After polymerizing, using heat to melt and remove the
coverslip. 5) Section and view them on TEM. So far, the surface of the
cells have been stuck to the coverslip after removal or glass pieces
stuck to the sample.
I am open to suggestions for any other strategy.
Thanks.
Eunice


------------------------------------------------------------------------
---




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 13:24:52 2005



From: Dorota Wadowska :      wadowska-at-upei.ca
Date: Mon, 14 Mar 2005 15:30:37 ADT
Subject: [Microscopy] TEM: Spurr and methanol

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello everybody,
Is there a reason why methanol should not be used as a
dehydrating agent for samples to be embedded in Spurr resin?
Thanks
Dorota


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 13:27:29 2005



From: JoAnn Buchanan :      redhair-at-stanford.edu
Date: Mon, 14 Mar 2005 11:36:28 -0800
Subject: [Microscopy] removing glass coverslips from cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Eunice. In our lab we routinely process hippocampal neurons grown on
glass coverslips for TEM. Since I often have to relocate an individual
neuron that has been imaged by LM, I need to have the entire coverslip
intact so that I can find the cells. I embed in a Chang monolayer mold
using Embed 812 resin. When the coverslips are polymerized, I file down the
edges of the embedded coverslips with a metal file and immerse them in
hydrofluoric acid. The acid is in small plastic beakers kept under the fume
hood.(always wear nitrile gloves) After the coverslips are dissolved
(takes about 20-30 minutes). Remove the the plastic wafer with plastic
forceps and rinse well in running water. You can dry the wafers in the
embedding oven. I then stain them with toludine blue to see the cells in
the light microscope(unless I have done immuno-staining, then I leave them
unstained) to locate the cell of interest. Mark it with a marking pin that
in attached to a microscope objective. You can then punch out the cell
with a leather hole punch and reattach to a blank stub using resin. Trim
down to region of interest.
This way you can get many blocks from 1 coverslip and get a particular cell
if you need to.
If you work under the hood and wear gloves and use plastic containers, you
won't be exposed to the acid. The hydrofluoric acid is available in a
special container equipped with a safety pouring spout (Fisher).This is
more reliable than the popping off method that can leave glass fragments
which are bad for you knife.
Good luck, JoAnn

Department of Molecular and Cellular Physiology
Stanford University School of Medicine
Stanford, CA 94305
650-723-5856



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 14:07:28 2005



From: Dusevich, Vladimir :      DusevichV-at-umkc.edu
Date: Mon, 14 Mar 2005 14:17:26 -0600
Subject: [Microscopy] Bone sectioning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear all,

I have a problem sectioning non-demineralized bone. I pick sections on a
carbon coated formvar film, and I am getting multiple wrinkles on areas
with
mineral: http://www.umkc.edu/dentistry/microscopy/index-f.htm
Areas with soft tissue (on the right of the second picture) are good,
no wrinkles.

Any comments?

Thank you,

Vladimir


Vladimir M. Dusevich, Ph.D.
Electron Microscope Lab Manager
3127 School of Dentistry
650 E. 25th Street
Kansas City, MO 64108-2784

Phone: (816) 235-2072
Fax: (816) 235-5524
Web: http://www.umkc.edu/dentistry/microscopy



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 14:22:39 2005



From: Sherwood, Margaret :      MSHERWOOD-at-PARTNERS.ORG
Date: Mon, 14 Mar 2005 15:32:32 -0500
Subject: [Microscopy] viaWWW: removing glass coverslips from cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Randy,

That's OK except that some cells will not attach to plastic and need to be grown
on glass (and vice verse).

Peggy

Peggy Sherwood
Lab Associate, Photopathology
Wellman Center for Photomedicine (W224)
Massachusetts General Hospital
55 Fruit Street
Boston, MA 02114
617-724-4839 (voice mail)
617-726-6983 (lab)
617-726-3192 (fax)
msherwood-at-partners.org



-----Original Message-----
} From: Tindall, Randy D. [mailto:TindallR-at-missouri.edu]
Sent: Monday, March 14, 2005 1:38 PM
To: by way of MicroscopyListserver
Cc: microscopy-at-microscopy.com

Eunice,

Another alternative, if someone hasn't already mentioned it, is to grow
the cells on Thermonox plastic coverslips and embed the coverslips
directly into your resin. You can section these like any other sample
and avoid having to remove a glass cover slip from resin. Thermonox
cover slips are designed for cell culture, should be readily available
the various scientific and EM supply houses and come in a variety of
sizes and shapes.

The drawback is that you will most likely have to section the
coverslips, and thus the cells, in cross-section, rather than having the
nice top-down (or bottoms-up, rather) view of the cell. If that's not a
problem, Thermonox is a great help.

Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility---We Do Small Well!
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.emc.missouri.edu









-----Original Message-----
} From: by way of MicroscopyListserver [mailto:echeung-at-eyetech.com]
Sent: Friday, March 11, 2005 11:03 PM
To: microscopy-at-ns.microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (echeung-at-eyetech.com) from
http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on
Friday, March 11, 2005 at 17:29:24
------------------------------------------------------------------------
---

Email: echeung-at-eyetech.com
Name: Eunice Cheung

Organization: eyetech

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi.
I was wondering if there were any available step by step protocols on
removing glass coverslips from cells grown on them. I have been trying
to study the en face structure of cells. 1) growing them on coverslips
2)Processing them with Spurrs for ultramicronomy 3) using BEEM capsules
filled with Spurrs resin and inverting them in cotact with the glass
coverslips 4) After polymerizing, using heat to melt and remove the
coverslip. 5) Section and view them on TEM. So far, the surface of the
cells have been stuck to the coverslip after removal or glass pieces
stuck to the sample.
I am open to suggestions for any other strategy.
Thanks.
Eunice


------------------------------------------------------------------------
---





From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 21:05:48 2005



From: mikeraj-at-streamyx.com (by way of MicroscopyListserver)
Date: Mon, 14 Mar 2005 21:15:49 -0600
Subject: [Microscopy] viaWWW: dislocation loop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (mikeraj-at-streamyx.com) from
http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on
Sunday, March 13, 2005 at 11:48:59
---------------------------------------------------------------------------

Email: mikeraj-at-streamyx.com
Name: Mike Marks

Organization: Cornell University

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi, I understand what a dislocation loop is, but what is a
prismatic dislocation loop ? Thanks !

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 14 21:06:01 2005



From: tbhatt-at-uic.edu (by way of MicroscopyListserver)
Date: Mon, 14 Mar 2005 21:16:11 -0600
Subject: [Microscopy] viaWWW: JB4 microtome

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (tbhatt-at-uic.edu) from
http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on
Monday, March 14, 2005 at 10:48:17
---------------------------------------------------------------------------

Email: tbhatt-at-uic.edu
Name: T. Bhattacharyya

Organization: OTO-HNS

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I am trying to get some info and preliminary training for
using JB4 microtome for plastic resin histology. Within the Chicago
area is there any such laboratory that may be helpful ?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 15 07:52:47 2005



From: John Shields :      jpshield-at-uga.edu
Date: Tue, 15 Mar 2005 09:02:47 -0500
Subject: [Microscopy] SEMS at Florida

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello everyone,
The Southeastern Microscopy Society (a LAS of MSA) will be
having its annual meeting May 18-20, 2005 at Pensacola
Beach, Florida.
Invited Speakers include Dr. Mark Farmer from the NSF on NSF
funding opportunities for Instrumentation, Dr. Sara Miller
from Duke Univ. on Lab Preparation for Bioterrorism
Surveillance, Dr. Sandeep Shah of NASA on Microscopy
Techniques in the Investigation of the Space Shuttle
Columbia Accident, and Dr. Thom Hopen of the ATFE on
Forensic Microscopy.

Also included are Tutorials and demonstrations by Microscopy-
related Corporate members such as AETOS, Nikon, Skyscan
MicroCT and Hitachi.

SEMS meetings are always informative, instructive, and
relaxed for a society of its size. We encourage student
participation through the Ruska competition which offers a
monetary award and waivers of registration, banquet fees and
provides rooms for the
participating students.

More information on SEMS, the meeting this May, the Ruska
competition and our newsletter are available at
www.semicroscopy.org

Looking forward to seeing you there!
John Shields
--
John Shields
E.M. Lab, 151 Barrow Hall
University of Georgia
Athens, GA 30602
706-542-4080

John P. Shields
Center for Ultrastructural Research
151 Barrow Hall
University of Georgia
Athens, GA 30602

706-542-4080


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 15 12:54:21 2005



From: Jon Mulholland :      jwm-at-stanford.edu
Date: Tue, 15 Mar 2005 11:04:11 -0800
Subject: [Microscopy] LSM510 IRM contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear All,
We would like to use our LSM510 confocal to do reflection
interference contrasting (which we have no experience with). However, we
are not sure about the equipment we would need for this imaging technique
and if it can be applied to the confocal. As I understand it, the
technique normally requires, when using Hg illumination, an illumination
source polarizer, antiflex objective (which would have the quarter-wave
retardation plate mounted on the lens) and also an analyzer. We would be
using laser excitation and so it will already be polarized and we can
obviously setup the scope for reflective microscopy, we also have a Zeiss
63X antiflex lens.
What we don't have is the analyzer and an understanding of where
it would be placed in the 510 light path so it could work. Does anyone have
experience using the technique on a confocal, is it possible?

Jon Mulholland
Cell Sciences Imaging Facility
Beckman Center, B050
Stanford University School of Medicine
Stanford, CA 94305-5301

voice 650-725-7532
fax 650-725-4951

http://taltos.stanford.edu




From MicroscopyL-request-at-ns.microscopy.com Tue Mar 15 13:01:52 2005



From: jrobson-at-rdg.boehringer-ingelheim.com
Date: Tue, 15 Mar 2005 14:11:40 -0500
Subject: [Microscopy] Dye for Acrylic Adhesive

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

List:

I am trying to locate a dye that will allow me to detect adhesive creep in
an acrylic adhesive (pressure type) label system. I know that metal stains,
OsO4 & RuO4, can be used on acrylic adhesives, however, we are looking for
other approaches to advance our research. If you have any thoughts or
suggestions I would appreciate the assistance. Thank you, jr

John A. Robson





From MicroscopyL-request-at-ns.microscopy.com Tue Mar 15 13:55:52 2005



From: Alan Nicholls :      nicholls-at-uic.edu
Date: Tue, 15 Mar 2005 14:08:18 -0600
Subject: [Microscopy] MMMS Meeting March 24th, 2005

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Local MSA, MAS affiliate meeting announcement

For further information and abstracts for the student poster competition go
to:-
http://www.northwestern.edu/bioimaging/MMMS%20Website/meetings.htm

Imaging Biomolecular Interactions
Midwest Microscopy and Microanalysis Society
Co-sponsor: Biological Imaging Facility, Northwestern University

Thursday, March 24, 2005

Northwestern University
Pancoe Life Sciences Building
Pancoe/ENH Auditorium
Evanston, IL 60208

Meeting Schedule
Morning Session
9:00 Registration
9:45 Welcome and introduction
10:00 Victoria Froelich, University of Texas Health Sciences Center San
Antonio: FRET, dectection of molecular interactions.
11:00 Kanya Rajangam, Northwestern University: Cells in self-assembling gels.
11:30 Carina Holmberg and Heather Brignull, Northwestern University:
Imaging-based characterization of aggregates in neurodegenerative disease
models.
12:00-12:45 Lunch
12:45 Business Meeting

Afternoon session
1:00 H. Peter Lu, Paci. c Northwest National Laboratory: Single-molecule
biophysics.
2:00 Debbie Klos, Northwestern University: Recognition of membrane protein
cargo by the Rsp5 ubiquitin ligase.
2:30 Victoria Wu, Northwestern University: The expansive tube, the role of
the septate junction in Drosophila tracheal development.
3:00 Reception and Student Poster Competition Exhibit and Judging.

RSVP Required
Please send RSVP via email, phone, or fax to:
Arvid Casler, MMMS Program Coordinator
c/o MAI
P. O. Box 394
Mundelein, IL 60060
phone/FAX: (847) 566-7716
email: arvid_casler-at-fmo.com


Admission: Free to MMMS Members
MMMS Membership: $10.00
MMMS membership available at registration

Driving directions to Northwestern University Evanston Campus available on
the Web.
From the north or northwest, via Interstate 88, Interstate 90 or
Interstate 190
(Note: these are the directions to follow if traveling from O'Hare
International Airport)
http://www.northwestern.edu/campus/directions/evanston-north-northwest.html
From the west, via Interstate 94
http://www.northwestern.edu/campus/directions/evanston-west.html
From the west or southwest, via Interstate 55 or Interstate 80
http://www.northwestern.edu/campus/directions/evanston-west-southwest.html
From the south or southeast, via Interstate 94, Interstate 90, Interstate
80 or Interstate 57
http://www.northwestern.edu/campus/directions/evanston-south-southeast.html

PARKING
Parking permits are $4 and are nonrefundable.
Parking can be challenging on the NU Evanston Campus especially after 9AM.
Spaces in selected lots are available on a first-come-first-served basis.



Alan W Nicholls, PhD
Newsletter Editor - M3S
Research Resources Centre - East (M/C 337)
Room 100 Science and Engineering South Building
The University of Illinois at Chicago
845 West Taylor St
Chicago IL 60607-7058

Office: Room 110
Tel: 312 996 1227
Fax: 312 996 8091
nicholls-at-uic.edu
http://www.northwestern.edu/bioimaging/MMMS%20Website/index.htm



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 15 15:11:44 2005



From: Elaine F. Schumacher :      eschumacher-at-mccrone.com
Date: Tue, 15 Mar 2005 15:19:06 -0600
Subject: [Microscopy] Dye for Acrylic Adhesive

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

John,

Would a fluorescent dye that's soluble in your adhesive, and
fluorescence microscopy be of help to you? We've had success with
fluorescent dyes to track components such as fillers and adhesives in
several projects here.

Regards,

Elaine

*********************************************************************
Elaine F.Schumacher
Senior Research Scientist
McCrone Associates, Inc.
850 Pasquinelli Drive
Westmont, IL 60559-5539 USA
630-887-7100 (tel)
630-887-7417 (fax)
E-mail: eschumacher-at-mccrone.com
Web Site: www.mccrone.com
*********************************************************************
This message and any attachments are solely for the
intended recipient. If you are not the intended recipient,
disclosure, copying, use or distribution of the information
included in this message is prohibited.
*********************************************************************

-----Original Message-----
} From: jrobson-at-rdg.boehringer-ingelheim.com
[mailto:jrobson-at-rdg.boehringer-ingelheim.com]
Sent: Tuesday, March 15, 2005 1:12 PM
To: Microscopy-at-MSA.Microscopy.Com

List:

I am trying to locate a dye that will allow me to detect adhesive creep
in
an acrylic adhesive (pressure type) label system. I know that metal
stains,
OsO4 & RuO4, can be used on acrylic adhesives, however, we are looking
for
other approaches to advance our research. If you have any thoughts or
suggestions I would appreciate the assistance. Thank you, jr

John A. Robson







From MicroscopyL-request-at-ns.microscopy.com Wed Mar 16 10:55:35 2005



From: Karl Garsha :      garsha-at-itg.uiuc.edu
Date: Wed, 16 Mar 2005 11:04:54 -0600
Subject: [Microscopy] Position Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Greetings,
The ITG microscopist search has officially opened. The announcement is
posted at http://www.itg.uiuc.edu/announcements/VisitMicro.htm. Glen
Fried (gfried-at-itg.uiuc.edu; Office Phone: (217) 333-5493 ) or myself
(contact info. below) can be contacted if more information is desired.
I've appended the details below:

Applicants are being sought for the position of Visiting Microscopist in
the Imaging Technology Group (ITG) at the Beckman Institute at the
University of Illinois at Urbana-Champaign. The ITG provides both a
visualization facility and microscopy facility for campus researchers
with interests ranging from physics to biology and the arts. The
Microscopy Suite provides a wide selection of electron, scanning probe,
and optical imaging capabilities. Optical microscopy capabilities
include laser scanning confocal microscopy, multiphoton microscopy,
widefield transmitted and fluorescence microscopy, stereomicroscopy,
computer-assisted stereology, near-field scanning microscopy,
reflected/transmitted light micro-spectroscopy, and near IR imaging.
Further information can be found at
http://www.itg.uiuc.edu/ms/equipment/. The responsibilities of the
Visiting Microscopist include, but are not limited to:**

* Training new users, and assisting existing users with confocal
microscopy.
* Maintaining and benchmarking the performance of the confocal
microscope and following up on service calls.
* Training and assisting users with fluorescence, stereology, and
bright field imaging.
* Acting as a secondary resource for training and assisting users on
the electron and scanning probe microscopes.
* Develop advanced imaging technologies utilizing the capabilities
present in the ITG.
* Assisting in the evaluation, procurement, installation, and
maintenance of new instruments.
* Trouble shooting and user assistance on ancillary equipment.
* Maintaining facility equipment.
* Other duties as assigned.

Qualifications: Bachelor’s degree, two-years experience in optical
microscopy, and excellent verbal and written communications skills are
required. A strong background in biology and/or materials
characterization, experience in a microscopy user facility, knowledge
and/or experience in electron and scanning-probe microscopy, and an
advanced degree are preferred.

This is a full-time, visiting academic professional position with
university benefits and with the possibility of transitioning to a
permanent position. The start date is as soon as possible following the
close of the search. Salary is commensurate with qualifications and
experience. In order to ensure full consideration, applications must be
received by April 22, 2005. Applicants may be interviewed before the
closing date; however, no hiring decisions will be made until after that
date. To apply please send a letter of interest, resume, three letters
of reference, and contact information to:

Lori Heil
Beckman Institute
405 North Mathews Avenue
Urbana, IL 61801
Phone: 217-244-0170
Fax: 217-244-6219
E-mail: lheil-at-uiuc.edu

*/ The University of Illinois is an Affirmative Action, Equal
Opportunity Employer./*



--
Karl Garsha
Light Microscopy Specialist
Imaging Technology Group
Beckman Institute for Advanced Science and Technology
University of Illinois at Urbana-Champaign
405 North Mathews Avenue
Urbana, IL 61801
Office: B650J
Phone: 217.244.6292
Fax: 217.244.6219
Mobile: 217.390.1874
www.itg.uiuc.edu



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 16 12:06:19 2005



From: Geoff Williams :      willi1gl-at-cmich.edu
Date: Wed, 16 Mar 2005 13:15:25 -0500
Subject: [Microscopy] Agfa DuoScan 2500 bulb Source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

After about 4 years of solid use our scanner bulb seems to be failing.
 However I’m coming up empty on replacement sourcing.  And there is still a
small bit of doubt if it is the transmitted lamp at all anyway.
 
I’ve disassembled it enough to evaluate that it is not a trivial matter to
replace the bulb but that it should be possible.
 
My question to the group is: has anyone replaced a transmitted lamp on this
scanner (Agfa DuoScan T2500)?
And (as it appears) assuming this is as unavailable as it appears, what is
the current scanner on the market that is able to generate equivalent TEM
negative scans?
 
I know scanners have been discussed ad-nauseum, but the requirements I’m
looking for are: available and current, (and hopefully less expensive than
the Agfa).  Is the front runner still the Epson Perfection 4870?
 
Thanks!


Geoff Williams
Microscopy Facility Supervisor
 
Central Michigan University Biology Department Microscopy Facility
http://www.cst.cmich.edu/centers/microscopy/





From MicroscopyL-request-at-ns.microscopy.com Wed Mar 16 14:18:27 2005



From: Nancy Smythe :      smythen-at-musc.edu
Date: Wed, 16 Mar 2005 15:27:45 -0500
Subject: [Microscopy] Re: scanner

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have been using the Epson Perfection 4870 for almost a year now and am very happy with the results... Much less expensive than the Agfa and faster. It has served us well.



Nancy Smythe
Department of Otolaryngology
Head and Neck Surgery
Medical University of South Carolina
843-792-8835
843-792-0368 Fax




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 16 14:56:23 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Wed, 16 Mar 2005 16:06:09 -0500
Subject: [Microscopy] Histology/sections not adhering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

A colleague of mine was asked to section and stain Axolotl testis
(amphibian testis can be cut such that all stages of spermatogenesis
are present in the section). She was given 2 sets of tissue: ones
fixed in Bouin's (Formalin, glacial acetic acid and picric acid),
the others were fixed in Flemmings solution ( chromic acid, glacial
acetic acid and osmium tetroxide). The sections from the Bouin's
fixed tissue cut nicely but no matter what she has tried, they slip
off of the slides either during the de-paraffinization or
re-hydration steps. She has tried super-frost Plus slides,
gelatin-chrom alum, and even gelatin-chrome alum plus poly-l-lysine.
She lets the sections dry on the slide warming tray for a minimum of
24 hours. These slides are desired for the medical students here,
so she needs to cut and stain 120-150 slides. Very hard to do if you
can't get the sections to stick! She is about to try the
Flemmings-fixed tissue, but we just can't figure out what's going on
with the Bouins-fixed stuff.

Any ideas?

TIA,
Lee
--
Lee Cohen-Gould
Electron & Optical Microscopy Facilities
Weill Medical College of Cornell U.
(212)746-6146
Rms A-105, LC-207


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 09:53:03 2005



From: Jan Factor :      jfactor-at-ns.purchase.edu
Date: Wed, 16 Mar 2005 16:06:09 -0500
Subject: [Microscopy] Histology/sections not adhering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Years ago I did extensive serial sectioning of Bouin's-fixed,
paraffin-embedded tissues. I had very good results (i.e., sections did
not fall off slides during processing) using a light coating of albumen
fixative on plain, alcohol-cleaned microslides. Good luck.
--Jan

---------------------------------------
Jan Robert Factor, Ph.D.
Professor of Biology
---------------------------------------
Natural Sciences
Purchase College
State University of New York
735 Anderson Hill Rd.
Purchase, NY 10577
USA
---------------------------------------
Office Tel: 914-251-6659
Office Fax: 914-251-6635
E-mail: jfactor-at-ns.purchase.edu
or- jan.factor-at-purchase.edu
---------------------------------------


-------- Original Message --------

A colleague of mine was asked to section and stain Axolotl testis
(amphibian testis can be cut such that all stages of spermatogenesis
are present in the section). She was given 2 sets of tissue: ones
fixed in Bouin's (Formalin, glacial acetic acid and picric acid),
the others were fixed in Flemmings solution ( chromic acid, glacial
acetic acid and osmium tetroxide). The sections from the Bouin's
fixed tissue cut nicely but no matter what she has tried, they slip
off of the slides either during the de-paraffinization or
re-hydration steps. She has tried super-frost Plus slides,
gelatin-chrom alum, and even gelatin-chrome alum plus poly-l-lysine.
She lets the sections dry on the slide warming tray for a minimum of
24 hours. These slides are desired for the medical students here,
so she needs to cut and stain 120-150 slides. Very hard to do if you
can't get the sections to stick! She is about to try the
Flemmings-fixed tissue, but we just can't figure out what's going on
with the Bouins-fixed stuff.

Any ideas?

TIA,
Lee
--
Lee Cohen-Gould
Electron & Optical Microscopy Facilities
Weill Medical College of Cornell U.
(212)746-6146
Rms A-105, LC-207


--




From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 10:37:27 2005



From: Pat Connelly :      psconnel-at-sas.upenn.edu
Date: Thu, 17 Mar 2005 11:34:28 -0500
Subject: [Microscopy] Re: Histology/sections not adhering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} A colleague of mine was asked to section and stain Axolotl testis
} (amphibian testis can be cut such that all stages of spermatogenesis
} are present in the section). She was given 2 sets of tissue: ones
} fixed in Bouin's (Formalin, glacial acetic acid and picric acid),
} the others were fixed in Flemmings solution ( chromic acid, glacial
} acetic acid and osmium tetroxide). The sections from the Bouin's
} fixed tissue cut nicely but no matter what she has tried, they slip
} off of the slides either during the de-paraffinization or
} re-hydration steps. She has tried super-frost Plus slides,
} gelatin-chrom alum, and even gelatin-chrome alum plus poly-l-lysine.
} She lets the sections dry on the slide warming tray for a minimum of
} 24 hours. These slides are desired for the medical students here,
} so she needs to cut and stain 120-150 slides. Very hard to do if
} you can't get the sections to stick! She is about to try the
} Flemmings-fixed tissue, but we just can't figure out
Lee Cohen-Gould lcgould-at-med.cornell.edu
====================
Lee,
Bouin's fixation makes the tissue very hard, if I remember correctly.
This could make a difference in the tissue's adherence to the slides.

As to making the tissue stick - I have a technique that I learned
many years back when
I was attempting to make LM serial sections of lung tissue for a
reconstruction study.
I had tried every subbing solution and slide cleaning technique that
I could find in all
the histology books that I had. I also tried slides from 3 or 4
different companies.
Many sections would disappear at different times during the staining procedure.

Then I remembered Dorothy, a very experienced histology tech who
worked in the same lab
as I had several years before. I had never seen loose a paraffin
section! I went to her lab to observe just what she did differently
than I did from the time she put her coat on the hook
until she took a pinch of gelatin out of a rather large brown bottle
and she just sprinkled it
on top of the cold water that was in the water bath that would be
used to float the sections
when it had warmed to the proper temperature.
That was it! A pinch of gelatin - no subbing - no washing of the slides -
and I did not loose another section.

I asked Dorothy where she had learned the trick but she could not remember.
She thought that everyone knew about the gelatin in the water bath.
I guess that it is similar to learning electron microscopy techniques
from a book
without spending some time watching or talking with someone who has
things working.

Hoping it works for those in need!
Pat Connelly psconnel-at-sas.upenn.edu
The University of Pennsylvania
Department of Biology
Philadelphia, PA 19104-6018
215-898-7145


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 12:23:34 2005



From: Geoff McAuliffe :      mcauliff-at-umdnj.edu
Date: Thu, 17 Mar 2005 13:34:06 -0800
Subject: [Microscopy] Re: Histology/sections not adhering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Lee:

I suspect that his/her slides were not clean before the adhesive was
applied. In my many years of experience, slides out of the box are not
really clean, they have a very thin film of ?? on them. This film can be
removed by soapy water or alcohol. When such 'unclean' slides are coated
with your favorite adhesive, the adhesive lies on the film. When the
slides hit alcohols, the film, the adhesive and the section all come
off. When I started in microtechnique slides were obviously 'unclean'
so we always cleaned them with alcohol before applying egg albumin. Now
I wash my slides (in a slide staining rack) in hot, soapy water, rinse
well, rinse in distilled and dry in a clean (not used for wax) oven .
Then I coat with subbing soln, poly-L-lysine or 'silane' as needed. I
almost never lose sections.
Also, I always allow the water to drain from under the section by
proping the slide up (almost vertical) against the side of the warming
plate for a minute or two before laying the slide flat. Sections are
much more likely to lie flat after that. Sections that are flat look
translucent, sections with air under them look opaque and will come off.
Other than the above I know of no good reason for Bouin-fixed
material to not stick to a slide. If the Bouin material won't stick I
suspect the Flemming material won't either.

Geoff

Leona Cohen-Gould wrote:

} ------------------------------------------------------------------------------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
}
} A colleague of mine was asked to section and stain Axolotl testis
} (amphibian testis can be cut such that all stages of spermatogenesis
} are present in the section). She was given 2 sets of tissue: ones
} fixed in Bouin's (Formalin, glacial acetic acid and picric acid), the
} others were fixed in Flemmings solution ( chromic acid, glacial acetic
} acid and osmium tetroxide). The sections from the Bouin's fixed
} tissue cut nicely but no matter what she has tried, they slip off of
} the slides either during the de-paraffinization or re-hydration
} steps. She has tried super-frost Plus slides, gelatin-chrom alum, and
} even gelatin-chrome alum plus poly-l-lysine. She lets the sections dry
} on the slide warming tray for a minimum of 24 hours. These slides
} are desired for the medical students here, so she needs to cut and
} stain 120-150 slides. Very hard to do if you can't get the sections
} to stick! She is about to try the Flemmings-fixed tissue, but we just
} can't figure out what's going on with the Bouins-fixed stuff.
}
} Any ideas?
}
} TIA,
} Lee


--
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************





From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:41:31 2005



From: kssim-at-mmu.edu.my (by way of MicroscopyListserver)
Date: Thu, 17 Mar 2005 13:51:44 -0600
Subject: [Microscopy] viaWWW: donate one SEM to my lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (kssim-at-mmu.edu.my) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, March 15, 2005 at 09:39:40
---------------------------------------------------------------------------

Email: kssim-at-mmu.edu.my
Name: kssim

Organization: mmu

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Dear All,

I am looking for someone who can donate one SEM to my lab at Malaysia. Please kindly let me know if you have the workable SEM that you wish to donate.

Thank you
Kok Swee, Sim


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:40:57 2005



From: Austin.Elliott-at-manchester.ac.uk (by way of MicroscopyListserver)
Date: Thu, 17 Mar 2005 13:51:07 -0600
Subject: [Microscopy] viaWWW: Low temperature stages/chambers for light microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (Austin.Elliott-at-manchester.ac.uk) from http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, March 15, 2005 at 06:15:58
---------------------------------------------------------------------------

Email: Austin.Elliott-at-manchester.ac.uk
Name: Austin Elliott

Organization: University of Manchester UK

Title-Subject: [Microscopy] [Filtered] Low temperature stages/chambers for light microscopy

Question: My first encouter with this forum...

Does anyone know any good COMMERCIALLY-AVAILABLE low-temperature stages or (perhaps better) "environmental chambers" to do low temperature light microscopy on inverted microscopes?

Brief B/G - I have been working for many years with live-cell fluorescence/light microscopy/imaging on inverted 'scopes - all at room temperature! I have been reading a lot of fascinating stuff about cell freezing/ cryopreservation. Microscopy has been used to investigate this area, and a bunch of systems to produce a temp-controlled cryochamber for cells on the stage have been described, but all are basically one-off designs requiring a fair bit of engineering. I was wondering if there was anything commercially available.

Thanks

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:41:44 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Thu, 17 Mar 2005 14:51:34 -0500
Subject: [Microscopy] Re: Histology/sections not adhering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thanks again to everyone who responded. The gist of the advice is to
clean your slides (we did), and try one of the following adhesives
until you find one that works:
-good old fashioned egg albumin
-gel-chrom alum
-0.1% aqueous polyethyleneimine
-poly-l-lysine
-silane

thanks for the suggestion of trying different cleaning methods
(alcohol, acid or bleach) for different results.
I'll pass all of this along to my colleague and will be glad that I'm
not the one who has to try the various permutations of all of this
(this time).

Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:42:50 2005



From: Pietra-philipp-at-cooperhealth.edu (by way of MicroscopyListserver)
Date: Thu, 17 Mar 2005 13:52:47 -0600
Subject: [Microscopy] viaWWW: Kodak technical pan film 6415 discontinued

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (Pietra-philipp-at-cooperhealth.edu) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, March 16, 2005 at 11:13:36
---------------------------------------------------------------------------

Email: Pietra-philipp-at-cooperhealth.edu
Name: Philipp Pietra

Organization: Cooper university hospital department of Pathology

Title-Subject: [Microscopy] [Filtered] MListserver:Kodak technical pan film 6415 discontinued

Question: Does any one have any recommendations for substitute film since Kodak has discontinued their technical pan film 6415. I am using Zeiss TEM109 with transfiberoptic imaging camera using the 120mm film size. Any one else with this set up your input is valuable to me. EMS and Kodak have recommended the T-max 100 as a replacement, but for all my playing around I can't get the contrast to what I need. I am doing routine pathology and routinely there is no money so a new scope with a fancy digital camera is out of the question.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:42:19 2005



From: lilia.zhahalyak-at-trincoll.edu (by way of MicroscopyListserver)
Date: Thu, 17 Mar 2005 13:52:17 -0600
Subject: [Microscopy] viaWWW: TEM lab manual for cell (ventral mesencephanol) monolayer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (lilia.zhahalyak-at-trincoll.edu) from http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, March 15, 2005 at 12:34:53
---------------------------------------------------------------------------

Email: lilia.zhahalyak-at-trincoll.edu
Name: lilia zhahalyak

Organization: Trinity college, hartford, ct

Title-Subject: [Microscopy] [Filtered] MListserver: TEM lab manual for cell (ventral mesencephanol) monolayer

Question: I am looking for a scientific reference ie lab manuals for TEM primary cell culture with detailed steps of the procedure. I have Elaine Hunter "Practical electron Microscopy" 2nd edition, but it is not enough material for monolayer, and the procedure is rather vague. I even contucted M.A. Hayat, who wrote several books on the microscopy methods, but I do not think that he has time to unswer me. Do you have any suggestions?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:43:14 2005



From: jcai-at-nanostellar.com (by way of MicroscopyListserver)
Date: Thu, 17 Mar 2005 13:53:27 -0600
Subject: [Microscopy] viaWWW: electron diffraction pattern to study nanoparticle size

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (jcai-at-nanostellar.com) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, March 16, 2005 at 16:35:23
---------------------------------------------------------------------------

Email: jcai-at-nanostellar.com
Name: Juan

Title-Subject: [Microscopy] [Filtered] electron diffraction pattern to study nanoparticle size

Question: Dear listers:



I am wondering is there a well-established method to analyze nano-particle size from the electron diffraction pattern. For example, in the polycrystalline materials, can the full width of the half maximum of a diffraction spot be converted to the particle size? Is the Sherrer equation applicable in the electron diffraction analysis?



Any information is appreciated. Thanks.



Juan


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:43:50 2005



From: Stacey.Andringa-at-uc.edu (by way of MicroscopyListserver)
Date: Thu, 17 Mar 2005 13:54:02 -0600
Subject: [Microscopy] viaWWW: ISEL technique to assay for apoptosis-related DNA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (Stacey.Andringa-at-uc.edu) from http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on Thursday, March 17, 2005 at 06:31:11
---------------------------------------------------------------------------

Email: Stacey.Andringa-at-uc.edu
Name: Stacey Andringa

Organization: University of Cincinnati

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: A colleague of mine has a question about the ISEL technique to assay for apoptosis-related DNA fragmentation. He is unable to find any "kits" for this technique. He is interested in the ISEL procedure because he was told that it is better than TUNEL when assaying epithelial tissue/cells. The tissues being analyzed are stratified squamous epithelia.

Is it indeed true that ISEL is preferable to TUNEL and
are there validated protocols out there for mouse back-skin and forestomach?

Thanks for any help.

Stacey Andringa






---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 13:44:23 2005



From: kweidenh-at-montefiore.org (by way of MicroscopyListserver)
Date: Thu, 17 Mar 2005 13:54:34 -0600
Subject: [Microscopy] viaWWW: uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (kweidenh-at-montefiore.org) from http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on Thursday, March 17, 2005 at 12:12:00
---------------------------------------------------------------------------

Email: kweidenh-at-montefiore.org
Name: Karen Weidenheim

Organization: Montefiore Medical Center/Albert Einstein Coll Med

Title-Subject: [Microscopy] [Filtered] "new" uranyl acetate problem

Question: Dear Colleagues, since 11/04, upon receiving a new lot of UA, we have not been able to obtain usable contrast on diagnostic tissue specimens. We had been using standard UA/lead citrate staining with no problems for the 7 years I have been running the lab. Since 11/04, we have tried multiple permutations of ethanol/water solutions, changed the water we use several times, changed staining times (several times) all to no avail. WE are working with EM sciences; while the situation is murky, it sounds like the formulation of the UA salt has been changed without asking persons that actually use it, at least for diagnostic purposes.
I can no longer do diagnostic em work on anything other than low power kidney bx because there is insufficient contrast to delineate the subcellular organelle structures that I need to see, i.e. mitochondria, thick and thin filaments, etc. I cannot even find the area of the grid that may be pathologic based on my review of the screening sections. I get eyestrain on the scope because I cannot see.
Does anyone else have the same problem with this "new" reagent? and if so what have you done about it.


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 14:59:54 2005



From: Karl Hagglund :      hagglundk1-at-nku.edu
Date: Thu, 17 Mar 2005 16:09:48 -0500
Subject: [Microscopy] Hitachi s-570 maintenance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Colleagues,

I am in the process of cleaning the column and apertures on an Hitachi
S-570 SEM. I have found most of the necessary tools, but cannot find
something called the fixed aperture take off tool. I have checked my
manual, and it lists two parts as aperture tools (tool for obj. aperture
exchange, and tool for obj. aperture set). I have photos of these, and
have both in my tool kit. Unfortunately, neither appears to be what is
pictured in the manual.

I have posted the diagram of the tool and its use at the following URL:

http://semlab.nku.edu/aperturetool.html

Does anyone out there have one of these that might be available?
Otherwise, if someone could tell us what it is called (or even a part
number). I cannot find anything called the fixed aperture take-off tool
in my parts list or manual. I am down to the last aperture on the
system, and have quite a stigmation problem that I am thinking is a
result of this one. I would love to get this microscope up and running
at its original potential again.

Any tips on how to rig something that would work in lieu of the tool
would be appreciated as well.

Karl Hagglund
Biological Sciences, SC-102B
Northern Kentucky University, Nunn Drive
Highland Heights, KY 41099
859-572-5238
http://semlab.nku.edu



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 17:50:39 2005



From: Allen R. Sampson :      ars-at-sem.com
Date: Thu, 17 Mar 2005 17:59:21 -0600
Subject: [Microscopy] RE: Hitachi s-570 maintenance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Look instead for a tube with a rod through the center from one end and a
fine internal thread on the other end. That final aperture is threaded on
the outside, so the tube is placed up against it, threaded on and pulled
down to release the holder. When putting it on, thread it on the tube,
push it in place and use the rod to apply pressure on it while unthreading
the tube.

The holder is just held in place by a compression ring and can easily be
removed without a tool. Strong fingernails can do it, or the wooden end of
a swab can be carefully placed in the opening, pushed sideways so that you
get some friction in the hole and then pulled down.

I haven't seen the tool you had a diagram of before. I've worked on a
number of 570s and they have always had the tube and rod tool I've tried to
describe.

Allen R. Sampson, Owner
Advanced Research Systems
317 North 4th. Street
Saint Charles, Illinois 60174
phone (630) 513-7093 fax (630) 513-7092
email mailto:ars-at-sem.com web www.sem.com


On Thursday, March 17, 2005 3:10 PM, Karl Hagglund
[SMTP:hagglundk1-at-nku.edu] wrote:
}
}
}
------------------------------------------------------------------------
------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
-------
}
} Colleagues,
}
} I am in the process of cleaning the column and apertures on an Hitachi
} S-570 SEM. I have found most of the necessary tools, but cannot find
} something called the fixed aperture take off tool. I have checked my
} manual, and it lists two parts as aperture tools (tool for obj. aperture
} exchange, and tool for obj. aperture set). I have photos of these, and
} have both in my tool kit. Unfortunately, neither appears to be what is
} pictured in the manual.
}
} I have posted the diagram of the tool and its use at the following URL:
}
} http://semlab.nku.edu/aperturetool.html
}
} Does anyone out there have one of these that might be available?
} Otherwise, if someone could tell us what it is called (or even a part
} number). I cannot find anything called the fixed aperture take-off tool
} in my parts list or manual. I am down to the last aperture on the
} system, and have quite a stigmation problem that I am thinking is a
} result of this one. I would love to get this microscope up and running
} at its original potential again.
}
} Any tips on how to rig something that would work in lieu of the tool
} would be appreciated as well.
}
} Karl Hagglund
} Biological Sciences, SC-102B
} Northern Kentucky University, Nunn Drive
} Highland Heights, KY 41099
} 859-572-5238
} http://semlab.nku.edu
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Thu Mar 17 18:13:08 2005



From: Lehman, Ann R :      Ann.Lehman-at-trincoll.edu
Date: Fri, 18 Mar 2005 09:30:00 -0500
Subject: [Microscopy] __ viaWWW: TEM lab manual for cell (ventral

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear List,
Here is my reply to Karl. I have removed the attachment from the Microscopy
Listserver copy.
MM
----- Original Message -----
} From: "Mary Mager" {mager-at-interchange.ubc.ca}
To: "Karl Hagglund" {hagglundk1-at-nku.edu}
Sent: Thursday, March 17, 2005 4:19 PM

Dear Lilia,

Why don't you drop by the EM Library (or see
http://www.trincoll.edu/~alehman/EM_Library.htm) where you could check
out the reference sources Trinity has on-site?? As I've told you, I
would be glad to help if we plan to set up some time together.

In summary, there are several widely-used routine techniques for dealing
with cultured cells. You need to be specific about whether you must
maintain a certain orientation. For example, do you want to examine the
points of attachment of the cells to their substrate, or would a random
sampling of cell orientations suffice? Also, what specifically do you
wish to study? Certain fixation protocols may be used to optimize
traditional morphology, others optimize preservation of immunologically
active sites, while still others can be used to track progressive
changes in cell constituents over time. As an alternative to standard
chemical-based protocols, you can use cryopreservation and/or
cryoultramicrotomy.

Good luck in your search. I don't doubt that someone on this List can
help you with your specific query. Otherwise, I am available during the
Spring Break next week, and my schedule opens up wide thereafter.

Prof L

Ann Hein Lehman
Assistant Director, Electron Microscopy Facility
Trinity College
300 Summit Street
Hartford, CT 06106
v. 860-297-4289
f. 860-297-2538
e. ann.lehman-at-trincoll.edu
w. http://www.trincoll.edu/~alehman


-----Original Message-----
} From: by way of MicroscopyListserver
[mailto:lilia.zhahalyak-at-trincoll.edu]
Sent: Thursday, March 17, 2005 2:52 PM
To: microscopy-at-microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (lilia.zhahalyak-at-trincoll.edu) from
http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on
Tuesday, March 15, 2005 at 12:34:53
------------------------------------------------------------------------
---

Email: lilia.zhahalyak-at-trincoll.edu
Name: lilia zhahalyak

Organization: Trinity college, hartford, ct

Title-Subject: [Microscopy] [Filtered] MListserver: TEM lab manual for
cell (ventral mesencephanol) monolayer

Question: I am looking for a scientific reference ie lab manuals for TEM
primary cell culture with detailed steps of the procedure. I have
Elaine Hunter "Practical electron Microscopy" 2nd edition, but it is not
enough material for monolayer, and the procedure is rather vague. I even
contucted M.A. Hayat, who wrote several books on the microscopy methods,
but I do not think that he has time to unswer me. Do you have any
suggestions?

------------------------------------------------------------------------
---




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 18 08:59:20 2005



From: Sherwood, Margaret :      MSHERWOOD-at-PARTNERS.ORG
Date: Fri, 18 Mar 2005 10:09:41 -0500
Subject: [Microscopy] Re: Histology/sections not adhering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

At one time, I used to use a product, Histostik, which was sold by Accurate
Chemical & Scientific Corp. (Westbury, NY). The procedure was as follows:

1. Agitate slides in 95% alcohol, then either wipe dry or drain dry.
2. Prepare Histostik solution as follows:
a. Add 6 drops of Histostik to 200ml of deionized or distilled water.
b. Shake and store in a brown bottle at 4 degrees C.
3. Immerse slide(s) into a staining dish containing Histostik solution for
3-5 seconds. Remove and allow slides to drain
dry.
4. Store coated Histostik slides in a dust free plastic slide box.

The procedure worked very well. Bottom line for any procedure: you need to
start with really clean slides.

Peggy

Peggy Sherwood
Lab Associate, Photopathology
Wellman Center for Photomedicine (W224)
Massachusetts General Hospital
55 Fruit Street
Boston, MA 02114
617-724-4839 (voice mail)
617-726-6983 (lab)
617-726-3192 (fax)
msherwood-at-partners.org



-----Original Message-----
} From: Leona Cohen-Gould [mailto:lcgould-at-med.cornell.edu]
Sent: Thursday, March 17, 2005 2:52 PM
To: microscopy-at-msa.microscopy.com

Thanks again to everyone who responded. The gist of the advice is to
clean your slides (we did), and try one of the following adhesives
until you find one that works:
-good old fashioned egg albumin
-gel-chrom alum
-0.1% aqueous polyethyleneimine
-poly-l-lysine
-silane

thanks for the suggestion of trying different cleaning methods
(alcohol, acid or bleach) for different results.
I'll pass all of this along to my colleague and will be glad that I'm
not the one who has to try the various permutations of all of this
(this time).

Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 18 09:46:18 2005



From: Walck, Scott D. :      walck-at-ppg.com
Date: Fri, 18 Mar 2005 10:56:08 -0500
Subject: [Microscopy] viaWWW: electron diffraction pattern to study

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

In my opinion, there are too many microscope parameters that will affect the width of the diffraction spots such as exposure, convergence angle, intermediate lens setting, condenser aperture for you to be able to do this with electron diffraction. However, the image mode of the TEM is available and it can be calibrated easily enough. (I recommend the Mag-I-Cal sample.) Why not use that. You do have to develop a method to properly disperse your samples on a grid.

-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
P. O. Box 11472 (letters)
Guys Run Rd. (packages)
Pittsburgh, PA 15238-0472
Walck-at-PPG.com
(412) 820-8651 (office)
(412) 820-8515 (fax)


-----Original Message-----
} From: by way of MicroscopyListserver [mailto:jcai-at-nanostellar.com]
Sent: Thursday, March 17, 2005 2:53 PM
To: microscopy-at-microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (jcai-at-nanostellar.com) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Wednesday, March 16, 2005 at 16:35:23
---------------------------------------------------------------------------

Email: jcai-at-nanostellar.com
Name: Juan

Title-Subject: [Microscopy] [Filtered] electron diffraction pattern to study nanoparticle size

Question: Dear listers:



I am wondering is there a well-established method to analyze nano-particle size from the electron diffraction pattern. For example, in the polycrystalline materials, can the full width of the half maximum of a diffraction spot be converted to the particle size? Is the Sherrer equation applicable in the electron diffraction analysis?



Any information is appreciated. Thanks.



Juan


---------------------------------------------------------------------------




From MicroscopyL-request-at-ns.microscopy.com Fri Mar 18 10:25:54 2005



From: Frank Macaluso :      macaluso-at-aecom.yu.edu
Date: Fri, 18 Mar 2005 11:34:35 -0500
Subject: [Microscopy] Re: viaWWW: uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Karen,
We have experienced the same understaining as you. We also obtain our UA
from EMS. After several unsuccessful attempts to get better staining by
increasing UA concentration, temperature, time etc..., we called Stacey
Kirsch. She told us that the UA was further depleted to be a safer product
for users. We discovered the new product was not staining as well as the
old. Stacey sent us two test samples that were dried differently to
increase staining ability (she also sent these test samples to other labs
for comparison). We found both were better, but preferred one over the
other. We are now using the new UA at an 8% aqueous solution for 20
minutes at room temperature followed by 3 minutes in lead citrate. Our old
staining protol called for 5% UA in 50% ethanol for 20 minutes at room
temperature. Stacey told me that she has been supplying this revised
formulation since February 1. It should be the same UA that you are
using. If you wish, we can give you some UA from our stock.

Recently someone posted a staining protocol on the listserver using
iso-butanol saturated water. It smells awful, but shows promise for
increased staining in our preliminary tests. You may want to try this:
I.M. Roberts, Journal of Microscopy
207:97 (2002).

I hope this helps,
Frank



} Email: kweidenh-at-montefiore.org
} Name: Karen Weidenheim
}
} Organization: Montefiore Medical Center/Albert Einstein Coll Med
}
} Title-Subject: [Microscopy] [Filtered] "new" uranyl acetate problem
}
} Question: Dear Colleagues, since 11/04, upon receiving a new lot of UA, we
} have not been able to obtain usable contrast on diagnostic tissue
} specimens. We had been using standard UA/lead citrate staining with no
} problems for the 7 years I have been running the lab. Since 11/04, we
} have tried multiple permutations of ethanol/water solutions, changed the
} water we use several times, changed staining times (several times) all to
} no avail. WE are working with EM sciences; while the situation is murky,
} it sounds like the formulation of the UA salt has been changed without
} asking persons that actually use it, at least for diagnostic purposes.
} I can no longer do diagnostic em work on anything other than low power
} kidney bx because there is insufficient contrast to delineate the
} subcellular organelle structures that I need to see, i.e. mitochondria,
} thick and thin filaments, etc. I cannot even find the area of the grid
} that may be pathologic based on my review of the screening sections. I
} get eyestrain on the scope because I cannot see.
} Does anyone else have the same problem with this "new" reagent? and if so
} what have you done about it.
}
}
} ---------------------------------------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Fri Mar 18 12:46:10 2005



From: Fred Pearson :      eoptics-at-mcmaster.ca
Date: Fri, 18 Mar 2005 13:57:14 -0500 (Eastern Standard Time)
Subject: [Microscopy] New Deadline for MSC2005 abstract submissions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Dear MSC-SCM members,

Please note that the date for abstract submissions to the May conference has been extended.
Please make sure you submit your abstracts as soon as possible but no later than
the 8th of April.

Please visit the conference website for submission details and to view the conference program.
We have planned a superb scientific program and several workshops. Make sure you do not miss
this opportunity to learn about new microscopy methods.

http://www.brockhouse.mcmaster.ca/MSC-SMC2005/

Looking forward to your conference contributions.

Marcia Reid (on behalf of organizing committee)


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 18 16:49:04 2005



From: pwebster-at-hei.org (by way of MicroscopyListserver)
Date: Fri, 18 Mar 2005 16:59:16 -0600
Subject: [Microscopy] viaWWW: Uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (pwebster-at-hei.org) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday, March 18, 2005 at 12:02:22
---------------------------------------------------------------------------

Email: pwebster-at-hei.org
Name: Paul Webster

Title-Subject: [Microscopy] [Filtered] Uranyl acetate problem

Question: To continue this interesting subject, which curiously relates to the subject Marc Pypaert brough up recently concerning uranyl acetate quality control:

When uranyl acetate was supplied in its undepleted form a saturated aqueous solution was around a 3% (w/v) concentration. More recently, in our lab we have only been able to get staining in some applications using a 5% solution. Now I read that 8% solutions are being recommended for usable staining.

I wonder what is being taken out of the uranyl acetate when it is being depleted?

On a related subject, the message I get from all the safety seminars I have to attend, is that the real poblem with uranyl acetate is not radioactivity but that it is toxic when ingested or inhaled. No amount of depleting will ever protect us from eating or inhaling the stuff.

Uranyl acetate disposal protocols take into account the toxic nature of the substance but becasue it is a natural isotope, there are no special protocols for disposing it as a radioactive substance.

Deep down I feel we were better off with the "old" radioactive uranyl acetate? Anyone know where I could buy a jar for myself? Maybe the "old" stuff is still being sold in less regulated parts of the world?

I welcome comments.

Paul Webster.



---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Mar 18 16:48:32 2005



From: david.audette-at-sylvania.com (by way of MicroscopyListserver)
Date: Fri, 18 Mar 2005 16:58:45 -0600
Subject: [Microscopy] viaWWW: ImageSlave board Manual

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (david.audette-at-sylvania.com) from http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on Friday, March 18, 2005 at 09:06:57
---------------------------------------------------------------------------

Email: david.audette-at-sylvania.com
Name: David Audette

Organization: OSRAM Sylvania

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hello,

I seemed to have recently acquired the last ImageSlave board in the market and was not lucky enough to also get a manual with it. I do have a some setup pages but the directions refer to an elusive userís manual. If anyone can help me locate this manual, I would appreciate that. I am trying to capture images from an AMRAY 1845FE.

TIA

Dave Audette
OSRAM SYLVANIA
71 Cherry Hill Drive
Beverly, MA 01915
david.audette-at-sylvania.com


---------------------------------------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Sat Mar 19 14:37:26 2005



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sat, 19 Mar 2005 15:47:21 -0500
Subject: [Microscopy] Swift microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Ritchie Sims wrote:
==============================================================
Does anyone know who, these days, supplies microscope stage point counter
attachments for petrographic use?

We have an old one made by Swift, of England, but my Google searching has
not turned up any current suppliers or manufacturers.
================================================================
There were two Swift microscope companies: one in the USA that is still
going and one in the Uk that was taken over by Prior Scientific some time
ago.

Most of the old Swift products are no longer available but you could try
contacting Prior to see of they can help.

E-mail: uksales-at-prior.com

Disclaimer: SPI Supplies has no commerical relationships with Swift or
Prior.

Chuck
===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: http://www.2spi.com
############################
==================================================




From MicroscopyL-request-at-ns.microscopy.com Sun Mar 20 05:20:56 2005



From: Barrie Wells :      Barrie.Wells-at-conwyvalley.com
Date: Sun, 20 Mar 2005 11:37:21 -0000
Subject: [Microscopy] Re: point Counter Supplier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ritchie Sims enquired about a point counter to fit a petrographic microscope
(i.e. a rotating stage).

Charles Garber suggested trying Prior instruments (who purchased Swift and
hence took over production of the J-0415C electro-mechanical stage).

I think that my company (Conwy Valley Systems) makes the only device that
fits this description. If you know of any other, I would be interested to
hear of it.

When Prior withdrew from this market in 2002, my company faced a dilemma,
because we have a software product (www.petrog.com) which depends on
existence of a stepping stage. We tried to persuade Prior to continue
manufacture but they were effectively defeated by the high quality of their
product: it was so reliable, they rarely sold replacements, and the market
was not growing sufficiently to justify continuing manufacture.
We therefore commissioned the electronics department at our local university
to design an updated stepper, fully automated in both x and y, which we now
manufacture and sell (in very small quantities).
If you are interested, you can find out more at www.SteppingStage.com
We are sure that this is the only device of its kind in the world; we would
not have gone to all the trouble and expense of creating it if there had
been any alternative.
If you have any suggestions for new markets for this device, we would be
very pleased to hear from you (and I apologise if that is a commercial
plug).

Regards,

Barrie Wells
Conwy Valley Systems Limited



From MicroscopyL-request-at-ns.microscopy.com Sun Mar 20 20:00:54 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Sun, 20 Mar 2005 18:18:00 -0800
Subject: [Microscopy] Re: viaWWW: Uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Paul
According Merck Index, UA (chemical salt, not unknown stuff) is dissolvable
in the water up to 10% (I don't remember the temperature, but believe it's
RT). As I mentioned before, from my prospective UA is uranyl-acetate:
acetic salt of uranium. Every NORMAL chemical company will indicate the
quality of the product: analytical grade etc. It also normally indicates
how much impurities present. For instance, 98% pure (something like
laboratory grade). All EM suppliers just redistribute UA (correct me if I
am wrong) and somehow (magically) keep forgetting to indicate how pure
their material is. If you want "good stuff", you need order it
from chemical company, not re-distributor. Merck is a great company and
perhaps best for inorganic compounds (no interest but happy user for
decades when was in Europe). The problem, how you bring stuff in US? I
also think that every EM supplier should have responsibility for the
product quality. Regarding "depleted" - uranium is uranium and 0.7%
presence (or absence in depleted) of 235U does not matter. You also right
that radioactivity of UA in not an issue: Americans distributed 120 metric
tons of depleted uranium (yes) over Kosovo (Yugoslavia) grounds during the
invasion. Depleted uranium is a component of bullets and artillery
rounds. There is about few tons (if I remember correctly) of depleted
uranium in every Boeing plane (civilian). Uranium present danger as any
(and every) heavy metal. It's extremely unhealthy to breath UA
powder. Personally, I am using UA from Ted Pella without any problem (no
interest, happy user for many years). Last time I purpose a few years ago,
so things may changed. Normally, I am using 3% stock solution. If there
is a problem with dissolving occurred, one should test quality of water -
slightly basic water (an indication of bacterial contamination) will reduce
solubility of UA. Aqueous UA solutions are stable at pH 5-5.4. Sergey


At 04:59 PM 3/18/2005 -0600, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 10:10:14 2005



From: Richard Edelmann :      edelmare-at-MUOhio.edu
Date: Mon, 21 Mar 2005 11:21:57 -0500
Subject: [Microscopy] LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

O.k., folks here is a weird one: I have a user who routinely uses LR
White, with good sucess, for emmbedding Arabidopsis (hypocotels).
However last week during an over night infiltration (as normal) 1-part ethanol
to 1-part LR White, the samples partially polymerized in the vials on a rotator.

Two questions:

(1) The LR White sample mix presently is a very viscous gel - i.e. not
completely polymerized. Any ideas for freeing/recovering the samples from
the gel? There are multiple samples in each vial and samples need to be flat
embedded for orientation.

(2) Any ideas what happened? I am leaning towards partially polymerized LR
White to start with - but maybe not. The LR White used was from a new bottle
well within expiration (Bad transportation or storage perhpas). However, the
LR White out of the bottle had normal viscosity (i.e. did not seem to be
partially polymerized before use). And how could it have polymerized 1:1 with
Ethanol, in sealed vials (tightly capped, wrapped in 3 layers of seal view & 3
layers of parafilm - perhaps overkill but it makes them happy i.e. the Ethanol
could not have evaporated).

The lab space is kept at 20-22C and certainly did not get over 30C. There
are no UV Lights in the hoods or room, and no windows (beside done over
night). No, likely polymerization source.


(No suppliers are being implicated here, and in fact it should be noted that
the supplier tech-support was very friendly to the user. We're just trying to
prevent a repeat, and having to start the samples all over again).


Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Director
350 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513.529.5712 Fax: 513.529.4243
E-mail: edelmare-at-muohio.edu
http://www.emf.muohio.edu

"RAM disk is NOT an installation procedure."


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 10:40:53 2005



From: Elaine Humphrey :      ech-at-interchange.ubc.ca
Date: Mon, 21 Mar 2005 08:53:27 -0800
Subject: [Microscopy] Re: Cost Recovery in Imaging Facilities

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Debby
As another response from Canada I would like to corroborate that what
you suggest is exactly right.

This is a facility where the only hard money coming in from the
University was my wages and everything else had to be cost recovery.
But a large number of the faculty are finding that their research
budgets are getting cut and are asking for help for the cost of
research internally. The University has responded by allocating some
indirect costs. It is still only soft money but it sure makes a
difference. It means that I can help keep the cost of research to the
individual PI at a reasonable level. Some of the faculty are also
helping write the federal grants for service contracts and personel.

We have done a good job in showing how easy it is to get a good TEM,
SEM and confocal image. Hence the graduate students own network plays
a part and they push to get the training. By making this facility
easy to use and welcoming, last year we had over 730 users from 258
account holders and 68 departments and groups. I have three full time
technicians now and a half time office person to help me and they are
all worth their weight in whatever metal is the most expensive at the
moment! But the key is the faculty. Without the need of the
research faculty there would be no need for this facility. As it is,
the need is still growing not diminishing around here. It is largely
up to the faculty to put the case to the administrators.
Administrators listen to faculty where they don't have to listen to
managers.

Elaine



} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
Dr. Elaine Humphrey
Director, BioImaging Facility
President, Microscopy Society of Canada
University of British Columbia
6270 University Blvd, mail-stop Botany
Vancouver, BC
CANADA, V6T 1Z4
Phone: 604-822-3354
FAX: 604-822-6089
e-mail: ech-at-interchange.ubc.ca
website: www.emlab.ubc.ca


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 12:16:08 2005



From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Mon, 21 Mar 2005 12:28:37 -0600
Subject: [Microscopy] LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Richard,

The exact same thing happened to me a couple years back, turning the LRW
into something resembling cottage cheese. Dr. Tom Phillips here at UM
correctly identified the source of the problem as too much residual
osmium in the tissue. We don't normally osmicate tissues destined for
immuno work, but a client had brought the tissue in already fixed, but
apparently not washed enough to get rid of all the excess osmium. The
sample was trashed. We never did figure out how to rescue it.

Hope this helps. If your tissue was not exposed to osmium, I'd love to
hear what the problem ended up being with it.

Cheers,
Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility---We Do Small Well!
W122 Veterinary Medicine
University of Missouri
Columbia, MO 65211
Tel: (573) 882-8304
Fax: (573) 884-5414
Email: tindallr-at-missouri.edu
Web: http://www.emc.missouri.edu







-----Original Message-----
} From: Richard Edelmann [mailto:edelmare-at-MUOhio.edu]
Sent: Monday, March 21, 2005 10:22 AM
To: microscopy-at-MSA.Microscopy.com

O.k., folks here is a weird one: I have a user who routinely
uses LR White, with good sucess, for emmbedding Arabidopsis
(hypocotels).
However last week during an over night infiltration (as normal) 1-part
ethanol to 1-part LR White, the samples partially polymerized in the
vials on a rotator.

Two questions:

(1) The LR White sample mix presently is a very viscous gel - i.e. not
completely polymerized. Any ideas for freeing/recovering the samples
from the gel? There are multiple samples in each vial and samples need
to be flat embedded for orientation.

(2) Any ideas what happened? I am leaning towards partially polymerized
LR White to start with - but maybe not. The LR White used was from a
new bottle well within expiration (Bad transportation or storage
perhpas). However, the LR White out of the bottle had normal viscosity
(i.e. did not seem to be partially polymerized before use). And how
could it have polymerized 1:1 with Ethanol, in sealed vials (tightly
capped, wrapped in 3 layers of seal view & 3 layers of parafilm -
perhaps overkill but it makes them happy i.e. the Ethanol could not have
evaporated).

The lab space is kept at 20-22C and certainly did not get over
30C. There are no UV Lights in the hoods or room, and no windows
(beside done over night). No, likely polymerization source.


(No suppliers are being implicated here, and in fact it should
be noted that the supplier tech-support was very friendly to the user.
We're just trying to prevent a repeat, and having to start the samples
all over again).


Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Director
350 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513.529.5712 Fax: 513.529.4243
E-mail: edelmare-at-muohio.edu
http://www.emf.muohio.edu

"RAM disk is NOT an installation procedure."




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 14:18:15 2005



From: Yang, Ann-Fook :      YANGA-at-agr.gc.ca
Date: Mon, 21 Mar 2005 15:24:35 -0500
Subject: [Microscopy] LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Hi Richard,

I had similar experience. There was nothing wrong when I was doing immunogold fixation and embedding but it gave me results as you described when I was embedding osmium-fixed tissue. I had been unable to rescue any of those tissue stuck in gooey LRW.

It happened to osmium fixed tissue and older LR White. The shelf-life of LR White is not long if it is catalyzed by the manufacturer or vendor. Your new bottle might have been sitting in a cold room for a while and I bet you were embedding osmium-fixed tissue.

Nowadays, I purchase uncatalyzed LR White. I mix half a bottle with half of catalyst a day or two just before I need it and use a smaller bottle to keep the other half to minimize oxidation.
After mixing, I further divide them into aliquots so that I warm up small quantity each time. The results have been satisfactory.
I hope this helps.

Ann Fook Yang
EM Unit/ Unite EM
Agriculture and Agri-Food Canada/Agriculture et Agroalimentaire Canada
Telephone/Téléphone: 613-759-1638
Facsimile/Télécopieur: 613-759-1701
960 Carling Ave/960 Boul Carling
Ottawa,Ontario/Ottawa, Ontario
Canada
K1A 0C6
yanga-at-agr.gc.ca

Agriculture and Agri-Food Canada - Agriculture et Agroalimentaire Canada

-----Original Message-----
} From: Richard Edelmann [mailto:edelmare-at-MUOhio.edu]
Sent: Monday, March 21, 2005 11:22 AM
To: microscopy-at-MSA.Microscopy.com

O.k., folks here is a weird one: I have a user who routinely uses LR
White, with good sucess, for emmbedding Arabidopsis (hypocotels).
However last week during an over night infiltration (as normal) 1-part ethanol
to 1-part LR White, the samples partially polymerized in the vials on a rotator.

Two questions:

(1) The LR White sample mix presently is a very viscous gel - i.e. not
completely polymerized. Any ideas for freeing/recovering the samples from
the gel? There are multiple samples in each vial and samples need to be flat
embedded for orientation.

(2) Any ideas what happened? I am leaning towards partially polymerized LR
White to start with - but maybe not. The LR White used was from a new bottle
well within expiration (Bad transportation or storage perhpas). However, the
LR White out of the bottle had normal viscosity (i.e. did not seem to be
partially polymerized before use). And how could it have polymerized 1:1 with
Ethanol, in sealed vials (tightly capped, wrapped in 3 layers of seal view & 3
layers of parafilm - perhaps overkill but it makes them happy i.e. the Ethanol
could not have evaporated).

The lab space is kept at 20-22C and certainly did not get over 30C. There
are no UV Lights in the hoods or room, and no windows (beside done over
night). No, likely polymerization source.


(No suppliers are being implicated here, and in fact it should be noted that
the supplier tech-support was very friendly to the user. We're just trying to
prevent a repeat, and having to start the samples all over again).


Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Director
350 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513.529.5712 Fax: 513.529.4243
E-mail: edelmare-at-muohio.edu
http://www.emf.muohio.edu

"RAM disk is NOT an installation procedure."




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 15:37:50 2005



From: Ladd Research :      ladres-at-worldnet.att.net
Date: Mon, 21 Mar 2005 16:49:47 -0500
Subject: [Microscopy] Re: SEM: Critical-Point Dryer Info. needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

}
} ----- Original Message -----
} From: "Ladd Research" {ladres-at-worldnet.att.net}
} To: "Microscopy Listserver" {Microscopy-at-microscopy.com}
} Cc: {john-at-microvisionlabs.com}
} Sent: Friday, February 25, 2005 4:44 PM
} Subject: [Microscopy] Re: SEM: Critical-Point Dryer Info. needed
}
}
} } The Bomar CPD was the forerunner of the Ladd CPD #28000 which was first
} } manufactured around 1970. The 1500 and the 28000 had a number of
} } similarities. There are still quite a few 28000s functioning so I'd
expect
} } the 1500 might be okay. They were well made. E-mail me your address and
} } we'll see if we have a manual for either CPD.
} }
} } John Arnott
} }
} } Ladd Research
} } 83 Holly Court
} } Williston, VT 05495
} }
} } On-line Catalog: http://www.laddresearch.com
} }
} } tel: 1-802-658-4961(anywhere) or 1-800-451-3406(US)
} } fax: 1-802-660-8859
} } e-mail: ladres-at-att.net
} }
} }
}
} --------------------------------------------------------------------------
} --
} } --
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
} --------------------------------------------------------------------------
} --
} } ---
} }
} } Yes, someone just emailed me and said that The Bomar Company went under
} } about 20 years ago. I talked to Ernest Fullem Co. and they said they
} } had also carried the Bomar CPDs back in the day but never had any at
} } their location. They would just have them sent to the customer when one
} } was ordered. So they have no manuals or such. That is likely the same
} } with Ladd as well.
} } Looks like there are a few manuals out there from the emails I have been
} } getting and I should be ok on that front. Thanks for everyone's
} } responses. You have been very helpful to me.
} }
} } John Knowles
} } MicroVision Labs, Inc.
} }
} }
} } -----Original Message-----
} } } From: john hoffpauir [mailto:hoffpajo-at-yahoo.com]
} } Sent: Thursday, February 24, 2005 5:55 PM
} } To: John Knowles; Microscopy-at-microscopy.com
} } Subject: [Microscopy] Re: SEM: Critical-Point Dryer Info. needed
} }
} }
} }
} } ------------------------------------------------------------------------
} } ------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} } http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } ------------------------------------------------------------------------
} } -------
} }
} } i have found through google a refence that used the
} } bomar cpd. ladd sold it years back.
} } john
} } --- John Knowles {john-at-microvisionlabs.com} wrote:
} }
} } }
} } }
} } }
} } ------------------------------------------------------------------------
} } ------
} } } The Microscopy ListServer -- Sponsor: The
} } } Microscopy Society of America
} } } To Subscribe/Unsubscribe --
} } } http://www.microscopy.com/MicroscopyListserver
} } } On-Line Help
} } }

Ladd Research
83 Holly Court
Williston, VT 05495

On-line Catalog: http://www.laddresearch.com

tel: 1-802-658-4961(anywhere) or 1-800-451-3406(US)
fax: 1-802-660-8859
e-mail: ladres-at-att.net





From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 15:48:04 2005



From: Ian Hallett :      IHallett-at-hortresearch.co.nz
Date: Tue, 22 Mar 2005 10:02:43 +1200
Subject: [Microscopy] Re: LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Richard

This issue has cropped up on occasion in the past. Premature polymerisation is not necessarily related to osmication of the tissue but is often related to older resin or resin that has had poor storage, a frequent problem was overheating during transport and also to some components in the tissue or extraneous media. Like many people we have for many years bought the uncatalysed resin and made up a bottle only when we need it. We have had no problem in the 10 or so years we have done this. I dredged
through the archives for a comment I made a number of years ago (1997 to be exact) and append a quote from Roy Gillett of London Resin.

"The reason this pre-polymerisation occurs only with tissue must be
something to do with a tissue constituent catalysing polymerisation.
Older resin is much more susceptible to this than fresh monomer because
of the significant polymer growth that will inevitably have occurred
in the monomer. The most likely 'endogenous catalyst' from
previous experience is likely to be an amine or peroxide moiety in
the tissue"

Ian

Ian Hallett
HortResearch
Mt Albert Research Centre, Private Bag 92 169
Auckland, New Zealand
Fax +64 9 815 4201
Telephone +64 9 815 4200
EMail ihallett-at-hortresearch.co.nz


} } } "Richard Edelmann" {edelmare-at-MUOhio.edu} 22/03/2005 4:21:57 a.m. } } }


------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

O.k., folks here is a weird one: I have a user who routinely uses LR
White, with good sucess, for emmbedding Arabidopsis (hypocotels).
However last week during an over night infiltration (as normal) 1-part ethanol
to 1-part LR White, the samples partially polymerized in the vials on a rotator.

Two questions:

(1) The LR White sample mix presently is a very viscous gel - i.e. not
completely polymerized. Any ideas for freeing/recovering the samples from
the gel? There are multiple samples in each vial and samples need to be flat
embedded for orientation.

(2) Any ideas what happened? I am leaning towards partially polymerized LR
White to start with - but maybe not. The LR White used was from a new bottle
well within expiration (Bad transportation or storage perhpas). However, the
LR White out of the bottle had normal viscosity (i.e. did not seem to be
partially polymerized before use). And how could it have polymerized 1:1 with
Ethanol, in sealed vials (tightly capped, wrapped in 3 layers of seal view & 3
layers of parafilm - perhaps overkill but it makes them happy i.e. the Ethanol
could not have evaporated).

The lab space is kept at 20-22C and certainly did not get over 30C. There
are no UV Lights in the hoods or room, and no windows (beside done over
night). No, likely polymerization source.


(No suppliers are being implicated here, and in fact it should be noted that
the supplier tech-support was very friendly to the user. We're just trying to
prevent a repeat, and having to start the samples all over again).


Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Director
350 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513.529.5712 Fax: 513.529.4243
E-mail: edelmare-at-muohio.edu
http://www.emf.muohio.edu

"RAM disk is NOT an installation procedure."


______________________________________________________

The contents of this e-mail are privileged and/or confidential to the
named recipient and are not to be used by any other person and/or
organisation. If you have received this e-mail in error, please notify
the sender and delete all material pertaining to this e-mail.
______________________________________________________


From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 17:26:02 2005



From: Ken Tiekotter :      tiekotte-at-up.edu
Date: Mon, 21 Mar 2005 15:38:57 -0800
Subject: [Microscopy] Re: LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Richard,

Having used LRW from over 20 years, I have found there are two conditions,
which result in a partially polymerized gel: 1) older LWR; 2) older LRW
mixture 1:1 with EtOH creates enough heat to cause premature polymerization.

I have taken similar tissue as yours, simply cut away as much of the gel
material as is convenient, placed in 100% LRW in the refrigerator overnight
as an additional precaution to thoroughly infiltrate the gel + tissue with
the new LRW: then the next day, embed as usual and polymerize.

I have actually kept a bottle of older LWR medium grade around in the
refrigerator so I can utilize this odd behavior of partially polymerization.
In the case of small specimens for which I want a vertical orientation in
the gel capsules, I use the following: simple place the specimen in gel
capsule, allow partial polymerization at room temperature, removed and trim
'block' in the correct orientation, re-infiltrate with fresh LRW, embed and
polymerize in the oven. May not be conventional, but it has been a great
solution of specimens like nematodes.

As Ann-Fook Yang suggested, purchase uncatalyzed LR White. I mix it up all
at once and kept it in the refrigerator. I also refrigerate my EtOH series.

Regards,
Ken
_______________________________________
Kenneth L. Tiekotter, Adjunct Professor
The University of Portland
Department of Biology
5000 N Willamette Blvd.
Portland, OR 97203 USA

Tel.: 503.493.8861



On 3/21/05 8:21 AM, "Richard Edelmann" {edelmare-at-MUOhio.edu} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} O.k., folks here is a weird one: I have a user who routinely uses LR
} White, with good sucess, for emmbedding Arabidopsis (hypocotels).
} However last week during an over night infiltration (as normal) 1-part ethanol
} to 1-part LR White, the samples partially polymerized in the vials on a
} rotator.
}
} Two questions:
}
} (1) The LR White sample mix presently is a very viscous gel - i.e. not
} completely polymerized. Any ideas for freeing/recovering the samples from
} the gel? There are multiple samples in each vial and samples need to be flat
} embedded for orientation.
}
} (2) Any ideas what happened? I am leaning towards partially polymerized LR
} White to start with - but maybe not. The LR White used was from a new bottle
} well within expiration (Bad transportation or storage perhpas). However, the
} LR White out of the bottle had normal viscosity (i.e. did not seem to be
} partially polymerized before use). And how could it have polymerized 1:1 with
} Ethanol, in sealed vials (tightly capped, wrapped in 3 layers of seal view & 3
} layers of parafilm - perhaps overkill but it makes them happy i.e. the Ethanol
} could not have evaporated).
}
} The lab space is kept at 20-22C and certainly did not get over 30C. There
} are no UV Lights in the hoods or room, and no windows (beside done over
} night). No, likely polymerization source.
}
}
} (No suppliers are being implicated here, and in fact it should be noted that
} the supplier tech-support was very friendly to the user. We're just trying
} to
} prevent a repeat, and having to start the samples all over again).
}
}
} Richard E. Edelmann, Ph.D.
} Electron Microscopy Facility Director
} 350 Pearson Hall
} Miami University, Oxford, OH 45056
} Ph: 513.529.5712 Fax: 513.529.4243
} E-mail: edelmare-at-muohio.edu
} http://www.emf.muohio.edu
}
} "RAM disk is NOT an installation procedure."



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 18:10:30 2005



From: moss-at-relia.net
Date: Mon, 21 Mar 2005 17:22:52 -0700 (MST)
Subject: [Microscopy] Re: LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} I have had this problem in the past. In my case it was the
microcetrfuge tubes that we used. The colored tubes would do this
consistantly, whereas the clear (not white) were fine. Something in the
coloring agent was acting to catalyse the LR White.

Bill McManus
Mt Ogden Scientific Services
moss-at-relia.net
801/334-6677
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} O.k., folks here is a weird one: I have a user who routinely uses LR
} White, with good sucess, for emmbedding Arabidopsis (hypocotels).
} However last week during an over night infiltration (as normal) 1-part
} ethanol
} to 1-part LR White, the samples partially polymerized in the vials on a
} rotator.
}
} Two questions:
}
} (1) The LR White sample mix presently is a very viscous gel - i.e. not
} completely polymerized. Any ideas for freeing/recovering the samples
} from
} the gel? There are multiple samples in each vial and samples need to be
} flat
} embedded for orientation.
}
} (2) Any ideas what happened? I am leaning towards partially polymerized
} LR
} White to start with - but maybe not. The LR White used was from a new
} bottle
} well within expiration (Bad transportation or storage perhpas). However,
} the
} LR White out of the bottle had normal viscosity (i.e. did not seem to be
} partially polymerized before use). And how could it have polymerized 1:1
} with
} Ethanol, in sealed vials (tightly capped, wrapped in 3 layers of seal view
} & 3
} layers of parafilm - perhaps overkill but it makes them happy i.e. the
} Ethanol
} could not have evaporated).
}
} The lab space is kept at 20-22C and certainly did not get over 30C.
} There
} are no UV Lights in the hoods or room, and no windows (beside done over
} night). No, likely polymerization source.
}
}
} (No suppliers are being implicated here, and in fact it should be noted
} that
} the supplier tech-support was very friendly to the user. We're just
} trying to
} prevent a repeat, and having to start the samples all over again).
}
}
} Richard E. Edelmann, Ph.D.
} Electron Microscopy Facility Director
} 350 Pearson Hall
} Miami University, Oxford, OH 45056
} Ph: 513.529.5712 Fax: 513.529.4243
} E-mail: edelmare-at-muohio.edu
} http://www.emf.muohio.edu
}
} "RAM disk is NOT an installation procedure."
}
}



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 19:11:19 2005



From: Beaurega :      beaurega-at-westol.com
Date: Mon, 21 Mar 2005 20:24:05 -0500
Subject: [Microscopy] Re: viaWWW: Uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Paul,

I was a wet analytical research chemist for many years before being called
back into microscopy to restart a 'mothballed' TEM lab. Now it's gone and
I was forced to retire early. Anyway, I agree with all the comments of
Sergey.

A rose is a rose is a rose. Depleted uranium is something like 8% U235 and
depleted it is 4-6% U235. It is still quite radioactive. In either case,
the radioactivity doesn't change the chemical reactions of DU.
If you hold your 'pancake' geiger counter or equivalent near DU, it will
not make single clicks like background cosmic rays do. It will emit a
continuous tone on the times one scale even through a glass bottle and its
surrounding metal can. I had access to pounds of the stuff gathered over
the years by a chemist that I worked with in that wet lab. The gamma
radiation was detectable 12 feet away through walls and locked metal
cabinets. So inhalling DU or UA is bad.

Unfortunately, depleted uranium carries with it the implication it is 100%
depleted. It is only 40-60% depleted.
For example, a depleted silver mine seems to indicate no silver is left in
the mine.

If there is a microscopy problem with UA being used as a chemical reagent,
then it is with the purity, contamination, assay, or some other similar
problem.

A possible solution:
It is well known that before atomic absorption (atomic spectroscopy) that
the determinations of sodium were done using uranyl acetate as a raw
material. The UA was converted to zinc UA. The sodium was determined
gravimetrically in a platinum dish as the sodium salt. So, you might find
some 'old' UA available in a lab where it is almost impossible to dispose
of any unused UA cheaply today. These AA or wet labs might have pound or a
quarter-pound bottle of UA 'stored' somewhere in their labs. The
manufacturers / sellers, as I recall, were Fisher Chemical, B&A, and
Mallinkrodt.

Walk through a wet lab and see if your geiger counter goes off. There is
no need to open cabinets or metal containers. If you get a hit, the
original container should be a thin-walled (rusted?) steel capped and taped
can with a shrink wrapped plastic seal on the DUA bottle cap.

Good hunting (I hope),

Paul

Sergy,

There is a lot of unused older DUA in the US. It's not all on tanks and
ammo as U°. I found out the hard way that it is a almost impossible to
cheaply dispose of or sell UNUSED UA. Legal issues came up, shipping
issues were raised, people ran for cover, and risks over transportation
accidents were raised. I was told nobody, not even the US government,
wanted the stuff. I feel this 'old sodium UA reagent' could and should be
used in small quantities for research or by hospitals to help people like
Paul.


------------------------
At 04:59 PM 3/18/05 -0600, pwebster-at-hei.org wrote:
}
}
} ---------------------------------------------------------------------------
---
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 20:02:45 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 21 Mar 2005 18:30:34 -0800
Subject: [Microscopy] Re: viaWWW: Uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Mar 18, 2005, at 2:59 PM, by way of MicroscopyListserver wrote:

} I wonder what is being taken out of the uranyl acetate when it is
} being depleted?
}
} On a related subject, the message I get from all the safety seminars I
} have to attend, is that the real poblem with uranyl acetate is not
} radioactivity but that it is toxic when ingested or inhaled. No amount
} of depleting will ever protect us from eating or inhaling the stuff.
}
} Uranyl acetate disposal protocols take into account the toxic nature
} of the substance but becasue it is a natural isotope, there are no
} special protocols for disposing it as a radioactive substance.
}
Dear Paul,
The fissionable isotope U-235 is removed from depleted U, and, since
this is commonly done by gas centrifugation of UF6, it is hard for me
to see how the act of depletion could change the staining properties of
UAc. My best guess is that there is still some UFx and UFxO mixed in
with the UO2, but, since I don't know how the UF6 is treated after the
centrifugation process, this is strictly a wild guess. It is certainly
true that U-238, being an alpha-particle emitter, poses no threat from
the radiation unless it is inhaled or ingested. I have not looked into
the chemical toxicity, but I expect it is not good for you, as are a
lot of heavy metals. I know that Ra and, especially, Pu are toxic. An
inhaled U-containing particle that is retained in the lung can emit
alpha-particles that will be close enough to a living cell to cause a
lot of ionization within that cell, so inhaling a particle in the
sub-um size range will potentially expose a cell to a massive amount of
damage and can lead to cancer, and the same applies to an ingested
particle if the U finds its way into the tissue (rather than simply
being eliminated). Finally, whether there are or are not special
protocols for disposing of U depends on the state you are living in.
Some states require very expensive procedures for disposal, and in
others one can simply wash U down the sink with large amounts of water,
so everyone should check with the appropriate safety office (there
could also be applicable local ordinances).
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 21:18:40 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 21 Mar 2005 19:46:30 -0800
Subject: [Microscopy] Re: Re: viaWWW: Uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Mar 21, 2005, at 5:24 PM, Beaurega wrote:

} If you hold your 'pancake' geiger counter or equivalent near DU, it
} will
} not make single clicks like background cosmic rays do. It will emit a
} continuous tone on the times one scale even through a glass bottle and
} its
} surrounding metal can. I had access to pounds of the stuff gathered
} over
} the years by a chemist that I worked with in that wet lab. The gamma
} radiation was detectable 12 feet away through walls and locked metal
} cabinets. So inhalling DU or UA is bad.
}
Dear Beauriga,
What one reads from UAc at the distances you mention is, indeed, gamma
radiation, which is emitted from some of the isotopes in the U decay
chains. There will be a (nearly) steady state established where each
isotope in each decay chain (different for U-238 and U-235) has the
appropriate number of atoms such that the activities of each isotope
are the same; i.e., for each isotope, one atom is produced by the
parent isotope(s) for every one that decays. It is the gammas that can
cause damage at long distances, but each alpha which enters a
cell--thus acting over distances shorter that its very short range,
less than the thickness of the dead layer of the skin--does much more
damage that the gammas.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Mar 21 23:29:05 2005



From: Coetzee, Mr S. H Physics Science :      COETZEES-at-mopipi.ub.bw
Date: Tue, 22 Mar 2005 07:46:44 +0200
Subject: [Microscopy] Re: LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Richard

We had a similar problem.
We went the other way.
Opened another bottle of LR white and ditch that bottle of ethanol to general Lab use. The samples were lost. Forced it to polymerise but they went brittle. For us the amount of LR white in the bottle left and a unhappy user the cost of a smiley user was worth more. The EMU staff did the next run as a service (but also to verify that the user did it all correct.)
Best of luck ;)

We purchase only unanalysed LR White and we store all our LR white in the fridge. The lab does get occasionally very hot when the air-conditioning fails. Above 30 degree is not uncommon. On the batch that did not polymerise we had a few concerns:

1) The catalyser reacted with the packaging and had a brownish colour, but 90% of the bottle was used without problems.

2) The bottle might have got a bit old, 6 months after the catalyser were added.

3) The 99.999% ethanol might have absorbed to much water from the atmosphere.

We have resulted in Ethanol from glass containers only. Never let a bottle get past 6 months and never use a catalyst if it looks suspicious. The problem did not reoccur yet. Touch wood, ouch...



Since some mail do get Lost, Bounces, etc Please send a duplicate/copy of all urgent mail to:

coetzeesh-at-yahoo.co.uk {mailto:coetzeesh-at-yahoo.co.uk}

Mr S. H. Coetzee
Electron Microscope Unit
University of Botswana
Private Bag 0022
Gaborone
Botswana
Phone : +267 355 2462/5222
Mobile : +267 718 36547
Fax : +267 318 5097
e-mail : coetzees-at-mopipi.ub.bw

-----Original Message-----
} From: moss-at-relia.net [mailto:moss-at-relia.net]
Sent: Tuesday, March 22, 2005 2:23 AM
To: edelmare-at-MUOhio.edu
Cc: microscopy-at-msa.microscopy.com

} I have had this problem in the past. In my case it was the
microcetrfuge tubes that we used. The colored tubes would do this
consistantly, whereas the clear (not white) were fine. Something in the
coloring agent was acting to catalyse the LR White.

Bill McManus
Mt Ogden Scientific Services
moss-at-relia.net
801/334-6677
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} O.k., folks here is a weird one: I have a user who routinely uses LR
} White, with good sucess, for emmbedding Arabidopsis (hypocotels).
} However last week during an over night infiltration (as normal) 1-part
} ethanol
} to 1-part LR White, the samples partially polymerized in the vials on a
} rotator.
}
} Two questions:
}
} (1) The LR White sample mix presently is a very viscous gel - i.e. not
} completely polymerized. Any ideas for freeing/recovering the samples
} from
} the gel? There are multiple samples in each vial and samples need to be
} flat
} embedded for orientation.
}
} (2) Any ideas what happened? I am leaning towards partially polymerized
} LR
} White to start with - but maybe not. The LR White used was from a new
} bottle
} well within expiration (Bad transportation or storage perhpas). However,
} the
} LR White out of the bottle had normal viscosity (i.e. did not seem to be
} partially polymerized before use). And how could it have polymerized 1:1
} with
} Ethanol, in sealed vials (tightly capped, wrapped in 3 layers of seal view
} & 3
} layers of parafilm - perhaps overkill but it makes them happy i.e. the
} Ethanol
} could not have evaporated).
}
} The lab space is kept at 20-22C and certainly did not get over 30C.
} There
} are no UV Lights in the hoods or room, and no windows (beside done over
} night). No, likely polymerization source.
}
}
} (No suppliers are being implicated here, and in fact it should be noted
} that
} the supplier tech-support was very friendly to the user. We're just
} trying to
} prevent a repeat, and having to start the samples all over again).
}
}
} Richard E. Edelmann, Ph.D.
} Electron Microscopy Facility Director
} 350 Pearson Hall
} Miami University, Oxford, OH 45056
} Ph: 513.529.5712 Fax: 513.529.4243
} E-mail: edelmare-at-muohio.edu
} http://www.emf.muohio.edu
}
} "RAM disk is NOT an installation procedure."
}
}





From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 06:07:55 2005



From: j.bilde-at-risoe.dk
Date: Tue, 22 Mar 2005 13:20:09 +0100
Subject: [Microscopy] viaWWW: dislocation loop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Mike,

If you know what a dislocation loop is, you will probably also know the definition of the Burgers vector of a dislocation. A prismatic dislocation loop is a loop whose Burgers vector does not lie in the plane of the loop.

A dislocation loop whose Burgers vector does lie in the plane of the loop can glide in this plane. A prismatic loop, on the other hand, can only move in the plane of the loop by climb. It may, however, be able to glide along a prism defined by the Burgers vector and the line vectors of the loop.

Best regards,
Jorgen.

{:::} --- {:::} --- {:::} --- {:::} --- {:::} --- {:::}

Joergen B. Bilde-Soerensen
Senior Research Scientist, Ph. D.
Materials Research Department
Risoe National Laboratory
DK-4000 Roskilde
Denmark

e-mail: j.bilde-at-risoe.dk
phone: +45 4677 5802 (direct)
phone: +45 4677 4677 (switchboard)
fax: +45 4677 5758
website: http://www.risoe.dk/afm/Personal/jqbi/jqbi.htm



-----Original Message-----
} From: by way of MicroscopyListserver [mailto:mikeraj-at-streamyx.com]
Sent: 15. marts 2005 04:16
To: microscopy-at-ns.microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (mikeraj-at-streamyx.com) from
http://www.microscopy.com/MicroscopyListserver/MLFormMail.html on
Sunday, March 13, 2005 at 11:48:59
---------------------------------------------------------------------------

Email: mikeraj-at-streamyx.com
Name: Mike Marks

Organization: Cornell University

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Hi, I understand what a dislocation loop is, but what is a
prismatic dislocation loop ? Thanks !

---------------------------------------------------------------------------




From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 08:07:46 2005



From: dmphase-at-udc.ernet.in (by way of MicroscopyListserver)
Date: Tue, 22 Mar 2005 08:20:43 -0600
Subject: [Microscopy] viaWWW: Jeol JSM 5600 SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (dmphase-at-udc.ernet.in) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Monday, March 21, 2005 at 22:41:06
---------------------------------------------------------------------------

Email: dmphase-at-udc.ernet.in
Name: D.M.Phase

Organization: Inter University Consortium, India

Title-Subject: [Microscopy] [Filtered] Jeol JSM 5600 SEM

Question: Whether anybody knows website on which service mannual for above SEM machine is available online?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 09:48:42 2005



From: Frank Eggert :      eggert-at-mikroanalytik.de
Date: Tue, 22 Mar 2005 17:04:16 +0100
Subject: [Microscopy] X-ray lines and data

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

For all who have interest in EDX and X-ray data:

The new version of MA-Table software (X-ray line table, spectra
simulation, calculation of overlap situations and detection limits in
EPMA, ...) is ready for WEB-download. A display of of mass absorption
coefficient curves for any element and a calculation of absorption for
given energy and material thicknesses are new.

See the other new features:
http://microanalyst.net/manual.html#link3
Go to download:
http://microanalyst.net/registr.phtml

Best regards

Frank Eggert

======================
www.microanalyst.net
======================



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 11:18:43 2005



From: Richard =?iso-8859-1?Q?L=E9veill=E9?= :      leveille.richard-at-uqam.ca
Date: Tue, 22 Mar 2005 12:23:25 -0500
Subject: [Microscopy] VP-SEM/ESEM (non-cryo) observations of ice

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear microscopists,
We would like to be able to observe minerals and microorganisms in situ in
natural ice samples (permafrost, ground ice, sea ice) without destroying,
drying or modifying the sample.
We have an Hitachi FE-VP-SEM equipped with a Peltier-type cold stage (-30
deg C) and have tried looking at a few samples without much success in
observing micron-sized bacteria either using the ESED detector or
backscattered e-detector. We can however make general observations of the
ice over varying periods of time depending on sample type and size, as well
as observing conditions (temp, pressure, AV...).

All the literature I have found on ice observations is done with cryo-SEM
at much lower temperatures. Could this be done with a variable pressure
SEM or ESEM with a cold stage? Obviously sublimation is an issue and will
likely negatively impact the imaging capabilities, but can we expect a good
enough signal to see bacterial cells (roughly 1-2 microns)?
Any ideas or suggestions would be appreciated.
Cheers,
-Richard

Richard Léveillé
Postdoctoral Fellow-Stagiaire Postdoctoral
Centre Geotop-UQAM-McGill
Université du Québec à Montréal
C.P. 8888, Succ. Centre-Ville
Montréal, QC H3C 3P8
514-987-3000 5662# ***New/Nouveau***
514-987-4080
514-987-3635 (fax)
{http://www.geotop.uqam.ca/index.php?option=content&task=view&id=86&Itemid=79} W {http://www.geotop.uqam.ca/index.php?option=content&task=view&id=86&Itemid=79} ebpage





From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 13:56:06 2005



From: Lehman, Ann R :      Ann.Lehman-at-trincoll.edu
Date: Tue, 22 Mar 2005 15:10:16 -0500
Subject: [Microscopy] old uranyl acetate available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Listers,

I have some old U.A. from Fisher (#U-4) and Baker (#4192) free to a good
home, if I can figure out the correct way to ship it. Ideally the taker
would be someone local, and can come pick it up.

I also have uranyl nitrate (Fisher U-7, Baker #4196, and Mallinckrodt
#8640), as well as zinc uranyl acetate (Fisher #U-11). Also some thorium
compounds. Most of these reagents are in unopened brown bottles
apparently kept in the dark (in tins, etc). I would dearly love to
unload it all!

I acquired this material from a local lab that was shutting down, and I
carried various boxes and gadgets and some very nice pieces back to my
lab without realizing how much UA (and variants) was in there.

If anyone is interested, just let me know.

cheers,
Ann

Ann Lehman
Assistant Director
Electron Microscopy Facility
Mailstop: LSC314
Trinity College
300 Summit Street
Hartford, CT 06106
v. 860-297-4289
f. 860-297-2538
e. ann.lehman-at-trincoll.edu
w. http://www.trincoll.edu/~alehman



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 15:06:56 2005



From: Salamon, Daniel :      Daniel.Salamon-at-nrc-cnrc.gc.ca
Date: Tue, 22 Mar 2005 16:20:00 -0500
Subject: [Microscopy] Sputtering system usage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all,

Our facility recently bought a Gatan PECS 682 Ion Beam Sputtering system. Currently, we are using it for high-resolution coating of samples for our field-emission SEM. It is not currently under heavy use, but accounting keeps asking me to come up with a cost structure for next year.

Can you offer any information on how you set up your charging scheme (included in SEM, per sample, per unit time, per target material)?

Also, I would really appreciate it if anybody can give me an idea about how long different targets last before needing replacement. I understand this depends on system usage, but any information would help. Are there any other operating expenses?


Thanks,

Daniel Salamon
Technical Officer, Electron Microscopy

National Institute for Nanotechnology, NRC
W6-081 ECERF Bldg
9107-116 Street
Edmonton, AB. T6G 2V4

Phone: Office (780) 492 8878
Lab (780) 492 8872
DocuFax: (780) 492 8632



From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 17:46:54 2005



From: Judy Murphy :      murphyjudy-at-comcast.net
Date: Tue, 22 Mar 2005 15:58:25 -0800
Subject: [Microscopy] Re: Re: LR White Problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We had the problem of LR White partial polymerization only when there
was EtOH in the mixture. In our case, the fixative wasn't an issue. We
never had it with our researchers but started to see the problem with
students. More than 20 yrs ago, we started with using an infiltration
series with EtOH like one does with Epoxy and propylene oxide i.e. (2
parts EtOH:1 part LRW; 1part EtOH:1 part LRW; 1 part EtOH: 2 parts LRW).
This was followed by 2 changes in pure LRW. The researchers didn't have
the problem, but a few of the students did. I came to believe that it
was because some of the EtOH was left in the mixture i.e. the students
weren't as careful as the researchers. At that point, we stopped using
an infiltration series and went from 100% EtOH to 3 changes in pure
LRW. We never saw the problem again.

Partially polymerized blocks:
We tried desperately to save a couple of the partially polymerized
blocks. First of all, they NEVER polymerized even after 2 yrs in the
oven. We did however try to reclaim some of them. We cut the tissues
out of the partially polymerized blocks and put them first in propylene
oxide and let them soak and then in epoxy. We did this with vacuum
infiltration of the epoxy. We could cut them, but they weren't the
best. These were hard to get samples however so we felt lucky to get
anything.

Good Luck
Judy


Judy Murphy, PhD
Microscopy, Imaging & Lab Design Consultant
Stockton, CA 95219
murphyjudy-at-comcast.net

moss-at-relia.net wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 17:49:38 2005



From: tttan-at-simtech.a-star.edu.sg (by way of MicroscopyListserver)
Date: Tue, 22 Mar 2005 18:02:41 -0600
Subject: [Microscopy] viaWWW: High voltage feedthrough for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (tttan-at-simtech.a-star.edu.sg) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, March 22, 2005 at 10:42:51
---------------------------------------------------------------------------

Email: tttan-at-simtech.a-star.edu.sg
Name: TT Tan

Organization: SIMTech

Title-Subject: [Microscopy] [Filtered] High voltage feedthrough for TEM

Question: Hello all. May seems a little strange but i would like to know who are the manufacturers that produce such high voltage feedthrough for TEM. I would like to get one for experimental purposes. Thanks in advance

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 18:57:47 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Tue, 22 Mar 2005 17:11:48 -0800
Subject: [Microscopy] Re: Re: Re: viaWWW: Uranyl acetate problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Geiger counter normally detects only beta-emission. Alpha-emission may not
be detected other than by scintillation counter. Gamma-irradiation (and
some beta) in uranium is mostly due the quick slowing down the
beta-particles in the material, I forgot German word for that. It's also
called the "secondary" radiation. Uranium itself do not emit gamma. There
is some accumulation of U byproduct as correctly mentioned Bill Tivol,
which may emit alpha/beta or gamma particles (if gamma-photon is considered
to be a "particle"). From the "radiation" point of view - both 235U and
238U are both radioactive (no gamma). Alpha particles could travel just a
part of millimeter in the air. Piece of paper will stop most of them. So,
they are not dangerous to human being unless somehow you get them inside
your body. Beta particles also do not travel much - 1/2 in of plexiglass
(of glass bottle) effectively cut them out. Gamma is naturally present
everywhere: X-ray, travel by airplane, CAT scan (a lot!) etc. Amount of
gamma emitted by U is just negligible in compare how much charcoal power
plant emits into the atmosphere every second. If you happens to live in
the luxurious house with granite facade (even counter top on the kitchen
will count) - you will have noticeable higher radioactive background
around. Granite happens to release tiny amount of radon-gas, which is
highly radioactive and really dangerous. Some particular fillers for
concrete do the same. So, gamma is everywhere. In laboratory practice most
common and dangerous is beta-emission. You need to understand that
exposure-dose rapidly decreased with the distance (by power of two), so you
need to keep space between you and radioactive source: tweezers, glovers
etc. Water effectively adsorbs beta/alpha particles, so aqueous solutions
are less dangerous than solids .... and so on. If one wants to see how
"dangerous" the common laboratory environment is - take the jar with KCl
from your laboratory shelf, pour about 50 grams of KCl into Petri Dish and
measure radioactivity by Geiger counter... you will detect noticeable
radioactivity of some natural potassium isotopes... By the way, to measure
gamma - you need to use special badges from Radiation Safety Office or
special attachments to your Geiger counter (you will know if you have it,
because it' a few $K). Normal/standard Geiger counter DO NOT detect
gamma! Sergey

At 07:46 PM 3/21/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From MicroscopyL-request-at-ns.microscopy.com Tue Mar 22 19:28:27 2005



From: michael shaffer :      michael-at-shaffer.net
Date: Wed, 23 Mar 2005 07:18:42 -0330
Subject: [Microscopy] choosing a digital camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


How high are you talking about for the voltage?
And what is it for/from?

cheers

rtch


Date sent: Tue, 22 Mar 2005 18:02:41 -0600
To: microscopy-at-microscopy.com
} From: tttan-at-simtech.a-star.edu.sg (by way of MicroscopyListserver)

Hello Ritchie,

I am looking at 120kV. Basically its for feeding High Voltage into high
vacuum environment.

Regards
Thiam Teck
Research Engineer
Precision Measurement Group
SIMTech


-----Original Message-----
} From: Ritchie Sims [mailto:r.sims-at-auckland.ac.nz]
Sent: Wednesday, 23 March, 2005 9:41 AM
To: Tan Thiam Teck, Dr; microscopy-at-microscopy.com

It would appear that modern digital cameras dedicated to photomicroscopy
have balanced dynamic observation with pixel count. However, our
applications for a new optical workstations do not anticipate dynamic
observation (e.r., moving critters). Have I missed a camera manufacturer
that does offer a high resolution model, which does not provide a high frame
rate? For example, for real time observation and digital display, I'd
anticipate we'd only require 8fps (approximately 1024x768)... but would like
to capture imagery at high resolution, e.g., 5Mp.

Testamonials and sales representatives should contact me off list ...

tia & cheerios ... shAf :o)
Avalon Peninsula, Newfoundland
www.micro-investigations.com (in progress)



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 23 05:05:28 2005



From: michael shaffer :      michael-at-shaffer.net
Date: Wed, 23 Mar 2005 07:48:32 -0330
Subject: [Microscopy] SEM: EMF radiation from mass spectrometer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I would like to best predict how a mass spectrometer will affect an SEM,
both of which to be installed in the near future. The MS manufacturer
responds with several measurements near the front of the instrument and
magnet that are near 0.1 - 0.5 milliTestlas (measured at ~10cm) ... and that
around the sides and rear, that measuremnts fall off to earth's natural
levels( ~0.05mT). The space for the SEM, 15M distant, will be evaluated for
EMF at approximately 0.5milliGauss (10,000mT = 1mG), however at specific
frequencies.

Should I not be concerned with the emission from MS instrument because it is
static?

tia & cheerios ... shAf (a MS newbie) :o)
Avalon Peninsula, Newfoundland
www.micro-investigations.com (in progress)



From MicroscopyL-request-at-ns.microscopy.com Wed Mar 23 07:13:01 2005



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Wed, 23 Mar 2005 14:21:39 +0100
Subject: [Microscopy] Re: viaWWW: High voltage feedthrough for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

I don't know if TEM manufacturer buy they insulator by these companies,
but here are a few in that domain which have some choice. You can contact
too some vaccum component manufacturer like Huntington, Thermo-VG or KJ
Lesker. They sell such feedthrough.

The following are companies which manufacter themself.

In north america, and more or less elsewere
Ceramaseal, a company from Ceram Tech
ISI, distributed by MDC vacuum components

In europe
Frialit (formerly Degussa)
Desmarquet


Hope it helps

J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr

On Tue, 22 Mar 2005, by way of MicroscopyListserver wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (tttan-at-simtech.a-star.edu.sg) from http://www.msa.microscopy.com/MicroscopyListserver/MLFormMail.html on Tuesday, March 22, 2005 at 10:42:51
} ---------------------------------------------------------------------------
}
} Email: tttan-at-simtech.a-star.edu.sg
} Name: TT Tan
}
} Organization: SIMTech
}
} Title-Subject: [Microscopy] [Filtered] High voltage feedthrough for TEM
}
} Question: Hello all. May seems a little strange but i would like to know who are the manufacturers that produce such high voltage feedthrough for TEM. I would like to get one for experimental purposes. Thanks in advance
}
} ---------------------------------------------------------------------------
}




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 23 07:31:20 2005



From: Peter Van Osta :      pvosta-at-maia-scientific.com
Date: Wed, 23 Mar 2005 14:39:11 +0100
Subject: [Microscopy] Re: choosing a digital camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,

When choosing a camera there are two main basic principles to take into
account. The first is the spatial dimensions of the chip and the second
is its dynamic range.

The size of the chip and the size of its pixels will determine its
spatial qualities (simplified). The larger the chip size the larger its
field of view can be. The smaller the individual pixel size the better
its spatial resolution at a given resolution of the objective
(http://ourworld.compuserve.com/homepages/pvosta/pcrnyq.htm). With small
pixel sizes you do not need to add intermediate magnification steps to
capture the full resolution of the objective.

However as a general rule of thumb, the dynamic range of the camera
decreases with decreasing pixel size. Cooling a CCD wil decrease the
contribution of its own "thermal noise" and increases its usefull
dynaimc range.

The more pixels the more data you need to transfer to a PC. For consumer
grade cameras, USB and even FireWire will do. For more bandwidth a
CameraLink camera is better, but as USB and FireWire now come with every
computer this is of course easier.

Regards,

Peter

----------------------------------------------
Peter Van Osta, MD

Director Imaging
MAIA SCIENTIFIC
Cipalstraat 3
B-2440 Geel, Belgium
Tel.: +32 (0)14 570 620
Mobile: +32 (0)497 228 725
Fax.: +32 (0)14 570 621
Email: pvosta-at-maia-scientific.com
Website: www.maia-scientific.com
A Harvard Bioscience Company
----------------------------------------------

============================================================
michael shaffer wrote:
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} It would appear that modern digital cameras dedicated to photomicroscopy
} have balanced dynamic observation with pixel count. However, our
} applications for a new optical workstations do not anticipate dynamic
} observation (e.r., moving critters). Have I missed a camera manufacturer
} that does offer a high resolution model, which does not provide a high frame
} rate? For example, for real time observation and digital display, I'd
} anticipate we'd only require 8fps (approximately 1024x768)... but would like
} to capture imagery at high resolution, e.g., 5Mp.
}
} Testamonials and sales representatives should contact me off list ...
}
} tia & cheerios ... shAf :o)
} Avalon Peninsula, Newfoundland
} www.micro-investigations.com (in progress)




From MicroscopyL-request-at-ns.microscopy.com Wed Mar 23 07:51:53 2005



From: Karl Hagglund :      hagglundk1-at-nku.edu
Date: Wed, 23 Mar 2005 09:05:06 -0500
Subject: [Microscopy] Thanks S-570 maintenance questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thanks to all on the list who responded to my questions about the S-570
objective aperture. I received answers and the correct illustration for
my manual within an hour or so of posting my note, and I am thankful for
each and every response.

Hitachi's service people were also in touch with me shortly after my
posting and I would like to especially thank them for helping me out
with a 20 year old instrument that is no longer under service contract.
Their knowledge and experience with the instrument helped us to
troubleshoot the ultimate source of the problem and obtain the proper
part numbers to fix it.


Karl Hagglund
Biological Sciences, SC-102B
Northern Kentucky University, Nunn Drive
Highland Heights, KY 41099
859-572-5238
http://semlab.nku.edu