From: Gerd Leitinger :      gerd.leitinger-at-meduni-graz.at
Date: Wed, 01 Jun 2005 14:33:59 +0200
Subject: [Microscopy] Re: viaWWW: osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Where do your cells grow?
I had problems with osmium fixation of cell cultures when they grew on
plastic membranes, and a nice lady from this list server suggested that
I should reduce the osmium tetraoxide with potassium ferrocyanide
because the osmium reacts with the plastic polymer.

We now grow the cells on "Aclar film" and get good membrane
contrast, with or without reducing the osmium tetraoxide. This film is
supposed to be chemically insensitive and can withstand dehydration
and embedding.
The film has to be pre-treated with poly L lysine to prevent the cells
from floating away, though.

yours

Gerd Leitinger


}
} Question: I have recently been having problems getting good
staining of the membranes of cultured cells with osmium. The cells
were fixed with 2% glutaraldehyde and post fixed in 1% osmium
tetroxide. I have tried 0.1M phosphate buffer, pH 7.4 and 0.1M
cacodylate buffer pH 7.3. When using these the membranes were not
visible at all. I also tried osmium with potassium ferrocyanide which
worked well with phosphate buffer but left a black precipitate. With
cacodylate buffer and potassium ferrocyanide the membrane definition
was a little better but still not adequate. If anyone has any suggestions
as to how to improve the membrane staining, they would be greatly
appreciated. Thank you.
}
} ---------------------------------------------------------------------------
}




Dr. Gerd Leitinger

Institut für Zellbiologie, Histologie und Embryologie
Medizinische Universität Graz
Harrachgasse 21
A-8010 Graz
Austria

Tel. ++43 316 380 4237
Fax. ++43 316 380 9625
Mailto: Gerd.Leitinger-at-meduni-graz.at




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 07:48:37 2005



From: Gerd Leitinger :      gerd.leitinger-at-meduni-graz.at
Date: Wed, 01 Jun 2005 14:44:31 +0200
Subject: [Microscopy] Methanol permeabilisation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear all,

I have a question that takes up the recent permeabilisation issue:

Is it possible to permeabilise membranes with methanol instead of
ethanol for pre-embedding immunoEM, and what effects does the
methanol have on the ultrastructure of membranes?

I have some experience with pre-embedding immunocytochemistry of
vibratome slices of nervous tissue. For this, I usually pre-treat the cells
with up to 50% ethanol for a few minutes, and then add 0.05% saponin
to all the antibody solutions, and usually get quite good staining in the
outer regions of the slices.

I now wanted to do a similar procedure on cultivated cells and have
found that the antibodies do not penetrate into the cells at all! For light
microscopic immunocytochemistry, I have found that the antibodies
only penetrate if these cells are pre-treated with 100% methanol. For
immunoEM, I only tested my "usual" protocol of 10, 30, and 50%
ethanol for a few minutes each. Does anybody know whether it is worth
trying methanol for immunoEM, and perhaps applying the methanol for
a longer time, perhaps 30 minutes?

thank you

Gerd Leitinger


Dr. Gerd Leitinger

Institut für Zellbiologie, Histologie und Embryologie
Medizinische Universität Graz
Harrachgasse 21
A-8010 Graz
Austria

Tel. ++43 316 380 4237
Fax. ++43 316 380 9625
Mailto: Gerd.Leitinger-at-meduni-graz.at




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 08:17:07 2005



From: Aleksandr Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Wed, 01 Jun 2005 14:16:37 +0100
Subject: [Microscopy] Re: Methanol permeabilisation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Gerd,

Ethanol and methanol will destroy all the membranes you have in the
sample. You will see only some blobs of different densities instead of
nicely preserved cells.
For pre-embedding EM on cultured cells people use saponin 0.1% or
digitonin (do not remember %), even Triton X-100 is too harsh.

Gerd Leitinger:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 08:52:01 2005



From: Philip Oshel :      peoshel-at-wisc.edu
Date: Wed, 01 Jun 2005 08:51:18 -0500
Subject: [Microscopy] Re: viaWWW: osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Robert,

Try adding 1% tannic acid to your glut and possibly the OsO4,
Mallinckrodt cat. #1764 seems to work best -- I believe because it's
a monomer, instead of various polymers.
This helps with the membrane preservation.
Is this for TEM or SEM?

Phil

} Email: rtemkin-at-mtsinai.on.ca
} Name: Robert Temkin
}
} Organization: Hospital for Sick Children Research Institute, Toronto, Canada
}
} Title-Subject: [Microscopy] [Filtered] osmium fixation of cultured cells
}
} Question: I have recently been having problems getting good staining
} of the membranes of cultured cells with osmium. The cells were fixed
} with 2% glutaraldehyde and post fixed in 1% osmium tetroxide. I have
} tried 0.1M phosphate buffer, pH 7.4 and 0.1M cacodylate buffer pH
} 7.3. When using these the membranes were not visible at all. I also
} tried osmium with potassium ferrocyanide which worked well with
} phosphate buffer but left a black precipitate. With cacodylate
} buffer and potassium ferrocyanide the membrane definition was a
} little better but still not adequate. If anyone has any suggestions
} as to how to improve the membrane staining, they would be greatly
} appreciated. Thank you.
}
} ---------------------------------------------------------------------------

--
Philip Oshel
Supervisor, BBPIC microscopy facility
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)
http://www.ansci.wisc.edu/microscopy.htm


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 10:10:03 2005



From: Gretchen.Ziegler-at-leica-microsystems.com
Date: Wed, 1 Jun 2005 10:09:02 -0500
Subject: [Microscopy] position open at West Chester University

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html





Hello All,

I am posting this as a favour to Dr. Helen Reid at West Chester University:

The university is looking for a chemical microscopist for the Fall 2005
semester. If you are looking, or know anyone who is looking, for this
type of position, please contact Helen directly at:

hreid-at-wcupa.edu

Thanks,

Gretchen




Gretchen Ziegler, Sales Manager, Educational Division
Leica Microsystems, Inc.
P.O. Box 151 - Ocean Grove, NJ 07756
Phone: 732-897-9506 - Fax: 847-236-3013
Voicemail: 1-800-248-0665 ext. 5131
www.DiscoverMicroscopy.com


_____________________________________________________________________
This e-mail has been scanned for viruses by MCI's Internet Managed Scanning Services - powered by MessageLabs. For further information visit http://www.mci.com


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 11:24:53 2005



From: Pat Connelly :      psconnel-at-sas.upenn.edu
Date: Wed, 1 Jun 2005 12:06:40 -0400
Subject: [Microscopy] Re: via WWW: osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Greetings,
} It is really something to see one's name in print, especially
} for being the person to go to for information on a
} particular technique! Thanks Paul.
}
} With tissue cultured cells I have found it critical to fix them as
} soon as possible after they come out of the incubator.
} DO NOT WASH them! Simply decant the growth
} medium then immediately and gently flood the cells with
} the fixative.
}
} The fixative that works well on many different types of cultured
} cells as well as pelleted material and small tissue pieces is as follows:
} 1% Glutaraldehyde+1% Osmium tetroxide+0.05 M Phosphate Buffer
} -at-6.2 pH on ice and kept dark. Usually 45 minutes is fine.
}
} Yes, this fixative will start to oxidize. To avoid this have all
} the components
} on ice; mix in the glutaraldehyde immediately before the fixative is needed
} and allow sufficient volume to fix both the cells and the small amount of
} protein that is left on the cells after decanting the medium.
}
} Dr. Tilney INSISTS that the Electron Microscopy Sciences
} 8% Glutaraldehyde in the 10 ml vials be used
} at all times for his actin studies and that it be kept in
} a scintillation vial and discarded one week after opening
} the vial. The only reason for using the EMS product is that
} we always get excellent results and he refuses to even try
} another company.
}
} With the phosphate buffer one needs to wash well with cold
} (best available) water at least 3 times over 20 - 30 min. time
} AND remember to rinse the entire vessel (top too)
} at least once before adding cold 1% UA in water
} overnight in the refrigerator to avoid the dreaded uranyl-phosphate
} crystals.There is no light in our refrigerator so I do not need to worry
} if it really goes out when the door is closed.
}
} If the cells are to be fixed in the flask/petri dish for face-on sections,
} an ethanol dehydration is used and Ladd's LX-112 as an epon
} substitute. These do not melt the plastic.
} All other cases or cells grown in "Pernanox" dishes
} are acetone dehydrated and any epon substitute can be used.
}
} Contact me off-line if more information is needed.
} If there are several similar questions I'll post them.
}
} If this fixation does not show what you wish with the membranes
} try using an objective apperature that is a size smaller than is
} usually used in the TEM.
}
} Pat Connelly psconnel-at-sas.upenn.edu
} Dept. of Biology, University of Pennsylvania
} Philadelphia, PA 19104-6018


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 12:12:10 2005



From: Dave Joswiak :      joswiak-at-astro.washington.edu
Date: Wed, 1 Jun 2005 10:11:28 -0700 (PDT)
Subject: [Microscopy] looking for datacube software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I would be interested in hearing from the microscopy community what 3rd
party software is available for looking at datasets obtained from spectrum
imaging. As many of you know, spectrum images are 3d datasets which
contain an EDX (or EELS) spectrum at each pixel from a STEM image. We
have a Tecnai F20 FEG with STEM/EDX/EELS and want to probe deeper into
extracting useful information from these datacubes; our software can
generate basic element maps but we want to purchase software that is more
flexible and comprehensive and perhaps interactive too. For instance, we
would like to be able to highlight a subset of pixels in an image or
element map and see these points appear in a ternary diagram or other type
of plot. We can generate a spreadsheet of quantified elements or element
ratios for each pixel in a STEM image and can export the spreadsheet.
Thanks for your input.

Dave Joswiak
Univ. of Washington
Dept. of Astronomy
Seattle, WA


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 13:23:25 2005



From: Greg Erdos :      gwe-at-ufl.edu
Date: Wed, 01 Jun 2005 14:22:41 -0400
Subject: [Microscopy] Fwd: (Fwd) hawaii conference and LIMS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am forwarding this on behalf of Mr. Goodblatt.. He would like to have an
informal session at M&M in Honolulu regarding items he has
mentioned. Please reply to him directly if you are interested.





} ------- Forwarded message follows -------
} From: Avrum Goodblatt {goodblat-at-mail.med.upenn.edu}
} To: gwe-at-ufl.edu
} Subject: [Microscopy] hawaii conference and LIMS
} Copies to: luellen-at-mail.med.upenn.edu, jonni S. Moore
} {moorej-at-mail.med.upenn.edu} ,
} qcyu-at-mail.med.upenn.edu
} Date sent: Thu, 26 May 2005 12:01:48 -0400
}
}
} Managing data today has become more challenging due to the very large
} datasets
} being produced. It is not uncommon for a facility to turn out a full DVD
} of data in one
} day. Providing online access and backup becomes more complicated, especially
} when one also has to protect instrumentation from viruses and other attacks.
}
} In addition, usage and billing records should be tied into the LIMS but
} frequently
} these are two completely different systems. With compliance much more on
} the radar
} today, and funds coming from a wider variety of sources, better
} record-keeping is
} essential
}
} I am IT coordinator for a group of 9 resource facilities called
} PathBioResource
} (including BioMedical Imaging) in Pathology and Laboratory Medicine at the
} University of Pennsylvania. I would like to share with others what we are
} doing and
} learn from others ways to approach these issues. We might also find ways
} to share
} some tool development together in some open source fashion, or educate
} vendors as
} to what is needed.
}
} I will be in Honolulu at the conference from Sunday evening through Wednesday
} evening.
}
} Avrum Goodblatt
} goodblat-at-mail.med.upenn.edu
} 215 573 0675
}
}
}
} ------- End of forwarded message ---------
} Avrum Goodblatt goodblat-at-mail.med.upenn.edu
} BMCRC 215 573 0675

Gregory W. Erdos Ph.D.
Assistant Director, Biotechnology Program
Scientific Director, Electron Microscopy
P.O. Box 118525
217 Carr Hall
University of Florida
Gainesville, FL 32611
gwe-at-ufl.edu
352-392-1295
fax- 352-846-0251



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 13:48:41 2005



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Wed, 01 Jun 2005 13:47:53 -0500
Subject: [Microscopy] Re: looking for datacube software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I hope you are already familiar with the LISPIX package available through
NIST (http://www.nist.gov/lispix/doc/contents.htm). It is free and should
allow you to play with your data as you please. I have not used it myself
so I cannot comment on it.

The other question will be the data format. I don't know that vendors
afford export into a generic LISPIX format in a similar way as EDS data can
be exported to MSA format. Hopefully such portability will be forthcoming.
You can take the matter up with your EDS or EELS vendor.

Warren

At 12:11 PM 06/01/05, you wrote:

} I would be interested in hearing from the microscopy community what 3rd
} party software is available for looking at datasets obtained from spectrum
} imaging. As many of you know, spectrum images are 3d datasets which
} contain an EDX (or EELS) spectrum at each pixel from a STEM image. We
} have a Tecnai F20 FEG with STEM/EDX/EELS and want to probe deeper into
} extracting useful information from these datacubes; our software can
} generate basic element maps but we want to purchase software that is more
} flexible and comprehensive and perhaps interactive too. For instance, we
} would like to be able to highlight a subset of pixels in an image or
} element map and see these points appear in a ternary diagram or other type
} of plot. We can generate a spreadsheet of quantified elements or element
} ratios for each pixel in a STEM image and can export the spreadsheet.
} Thanks for your input.
}
} Dave Joswiak
} Univ. of Washington
} Dept. of Astronomy
} Seattle, WA

-------------------------------------------
No files should be attached to this message
-------------------------------------------
Warren E. Straszheim, Ph.D.
Materials Analysis and Research Lab
Iowa State University
46 Town Engineering
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
Web: www.marl.iastate.edu

Scanning electron microscopy, x-ray analysis, and image analysis of materials
Computer applications and networking



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 13:57:51 2005



From: Christine Weaver, McRI :      cweaver-at-mcri.org
Date: Wed, 1 Jun 2005 13:57:00 -0500
Subject: [Microscopy] Inter/Micro 2005 Preliminary Program

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Inter/Micro 2005 Preliminary Program
July 11-15, 2005
Talbott Hotel & McCrone Research Institute
Chicago, Illinois

McCrone Research Institute (McRI) cordially invites you to join us at
Inter/Micro 2005, an internationally recognized professional meeting
dedicated to applied microscopy, held each year in downtown Chicago.

The Inter/Micro 2005 conference will span 5 days. The first 3 days of the
conference will consist of a symposium, to be held at the Talbott Hotel,
featuring daily 15-20 minute paper presentations, a Monday evening session
with Brian J. Ford, Exhibitors (Tuesday and Wednesday), and McRI’s joint
meeting with the State Microscopical Society of Illinois including a
Banquet, Awards Ceremony, and Auction (Wednesday Evening). The symposium
will be followed by a special 2-day workshop (July 14th and 15th), to be
held at the laboratories and classrooms of McCrone Research Institute,
entitled "Introduction to Techniques of Forensic Soil Comparison" conducted
by Skip Palenik. Exhibitors and sponsor’s for Inter/Micro 2005 include:
Leica Microsystems, Educational Products, McCrone Research Institute, Bruker
Optics, Lomo America, State Microscopical Society of Illinois, Olympus
America, and Tienta Sciences.

To exhibit or sponsor (deadline July 1, 2005) please visit:
http://mcri.org/ExhibitorINFOandCONTRACT.pdf

The papers presented at Inter/Micro 2005 will be published in the
International Journal on microscopy, The MICROSCOPE. For more information,
call 312-842-7100 visit www.mcri.org or contact intermicro-at-mcri.org.


Monday, July 11
Techniques and Instrumentation

8:00 a.m. Registration: 1st floor, Gallery
9:00 a.m. Symposium: 1st floor, Delaware Room

Laughlin, Gary – McCrone Research Institute (McRI), Welcome

Havics, Tony – QEPI, A Menagerie of Microscopy

Weaver, Robert – McRI, Techniques of Controlling Depth of Field

Hinsch, Jan – Leica Microsystems, Modes of Viewing the Microscope’s Back
Focal Plane

Weaver, Robert – McRI, 40 Years of the Microscope Tricks of the Trade

Ford, Brian J. – Gonville & Caius College, University of Cambridge, A
Century of Brownian Motion

12:00 p.m. – 2:00 p.m. Lunch

Hinsch, Jan – Leica Microsystems, The Flatbed Scanner: Useful Beyond
Document Work

Erdman, Natasha – JEOL USA, Preparation of Clean Cross Sections Using JEOL
Cross Sectional Polisher (CP)

Bayard, Michael – Bayard Development, Microscopy of Solid Solution,
Co-Crystallized and Multi-Component Systems

Barnes, C.G. – Dow Chemical Company, Selective Etching of Nylon in Nylon/sPS
Fibers for Determination of sPS Morphology

Palenik, Mark – Microtrace, Characterization of Soot and Carbon Black by GC
and TEM

Sparenga, Sebastian – McRI, Identification of Abrasives in Smokeless Tobacco

8:00 p.m. Brian J. Ford, “Evening with Brian – Talk the Talk”


Tuesday, July 12
Environmental and Industrial Microscopy

8:00 a.m. Registration: 1st floor, Gallery
9:00 a.m. Symposium: 1st floor, Delaware Room

Bucher, Edith – Laboratorio Biologico - APPA Bolzano, Applied Aerobiology
and Melissopalynology in South Tyrol (Northern Italy) - Realization of a
Pollen Atlas

Chun-I Lee, James – University of Strathclyde, Diatom Distribution in
Danshuei River

Laughlin, Gary – McRI, Indoor Air Quality: Microscopy of House Dust
Particles

Speir, Jacqueline – McRI, Can Dispersion Aid in Amphibole Differentiation?

Chatfield, Eric – Chatfield Technical Consulting, TBA

Havics, Tony – QEPI, Asbestos: To Be a Fiber or Not To Be

12:00 p.m. – 2:00 p.m. Lunch

Ford, Brian J. – Gonville & Caius College, University of Cambridge, The
Discovery of Giardia

Alden, Harry – Smithsonian Center for Materials Research and Education, Hair
of the Dog, or How Blanket Fibers Can Get Your Goat

Welsh, Frank– Welsh Color and Conservation, Inc., Historic Paint Colors of
Spanish St. Augustine

Hills, Linda – CTL Group, From Rocks to Bridges: An Introduction to Cement
and Concrete Microstructure

Hagni, Ann – CTL Group, Failure Analyses of Construction Materials Using
Optical Microscopy

Rantanen, Walter – Integrated Paper Services, Inc., Matching Matches (Part
2)

5:00 p.m. – 6:00 p.m. Cocktail Hour with Exhibitors: 2nd floor, Victoria
Room

Wednesday, July 13
Chemical and Forensic Microscopy

8:00 a.m. Registration: 1st floor, Gallery
9:00 a.m. Symposium, 1st Floor Delaware Room

Crowe, John – FDA Forensic Chemistry Center, The Characterization of
Microchemical Test Resultant Crystalline Formations using PLM, FT-IR and
Raman Spectroscopy

Bowen, Andrew – Stoney Forensic, Inc., Recent Contributions of Chemical
Microscopy to the Analysis of Unknowns

Rothhaar, Katia – Tienta Sciences, Use of Drop Coated Deposition Raman for
Detection of Explosives

Hollifield, Jeff – Micro Analytical, Flow Chart for Rapid Identification of
Inorganic Compounds Using PLM

Ford, Brian – Gonville & Caius College, University of Cambridge, World's
Worst Microscopy

Palenik, Skip – Microtrace, TBA

12:00 p.m. – 2:00 p.m. Lunch

Reffner, John – Smith’s Detection, TBA

Boltin, Randy – MVA Scientific Consultants, A Class in Forensic Microscopy

Lawrence, Gene – San Diego County Sheriff's Crime Lab, Marine Coatings -
Components and Analysis

Grant, Ricky – 41st CST (WMD) US ARMY, Active National Guard, Weapons Of
Mass Destruction Terrorism Experts and Forensic Microscopy

Hopen, Thomas – ATF Forensic Science Laboratory, PLM Characterization of
Inorganic Constituents in Automobile Body Fillers

Hietpas, Jack – Microtrace, Microscopy of Explosives

Diaczuk, Peter – John Jay College of Criminal Justice, Determination of
Entry vs. Exit Bullet Holes in Garments using Light Microscopy

6:00 p.m. SMSI Banquet and Auction, Victoria Room
Dr. Osamu Shimomura, SMSI 2005 Émile Chamot
Award Recipient

Regards,
Christine

Christine Weaver
Inter/Micro Coordinator
McCrone Research Institute
2820 S. Michigan Avenue
Chicago IL 60616

312-842-7100
cweaver-at-mcri.org
http://www.mcri.org





From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 15:01:35 2005



From: Tea Meulia :      meulia.1-at-osu.edu
Date: Wed, 1 Jun 2005 16:02:43 -0400
Subject: [Microscopy] intestine em immunolocalizations

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We would like to do some em immunolocalizations on pig intestinal
tissue to detect virus infected cells. Can anyone direct me to some
literature? Which would be the best way to start pre- or
post-embedding labeling?

The antibodies, we will be using, work well in confocal microscopy
on whole tissue with Triton permeabilization, or on paraffin sections

Thank you.

Tea Meulia




--
***************************************
Tea Meulia, PhD
Director
Molecular and Cellular Imaging Center
Ohio State University/OARDC
1680 Madison Ave.
Wooster OH 44691

tel.: 330-263-3836 or -3828
fax: 330-202-3563

http://www.oardc.ohio-state.edu/mcic

*****************************************


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 15:17:01 2005



From: Elaine Humphrey :      ech-at-interchange.ubc.ca
Date: Wed, 01 Jun 2005 13:16:14 -0700
Subject: [Microscopy] gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello everyone
I have just had a request for the use of a somatoscope developed by
Gaston Naessens as they want colour at 30,000 times.

I was wondering if someone was pulling my leg but when I did a google
search I got
http://www.sumeria.net/tech/naessens.html
and I am still wondering if someone has a huge hoax going. Has anyone
got experience with this type of microscope?

Elaine


--
Dr. Elaine Humphrey
Director, BioImaging Facility
President, Microscopy Society of Canada (2003-2005)
University of British Columbia
6270 University Blvd, mail-stop Botany
Vancouver, BC
CANADA, V6T 1Z4
Phone: 604-822-3354
FAX: 604-822-6089
e-mail: ech-at-interchange.ubc.ca
website: www.emlab.ubc.ca


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 16:22:14 2005



From: rtemkin-at-mtsinai.on.ca (by way of MicroscopyListserver)
Date: Wed, 1 Jun 2005 16:21:32 -0500
Subject: [Microscopy] viaWWW: osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (rtemkin-at-mtsinai.on.ca) from http://www.microscopy.org/MicroscopyListserver/MLFormMail.html on Wednesday, June 1, 2005 at 08:06:18
---------------------------------------------------------------------------

Email: rtemkin-at-mtsinai.on.ca
Name: Robert Temkin

Organization: Hospital for Sick Children Research Institute, Toronto, Canada

Title-Subject: [Microscopy] [Filtered] osmium fixation of cultured cells

Question: Thanks for the responses but I guess I did leave out some details. I'm processing monolayers grown on glass coverslips. I do use en bloc staining with uranyl acetate. I'm using 2% aqueous UA for 45-60 minutes at room temperature. I use similar times for the glutaraldehyde and osmium fixations. I dehydrate with ethanol and embed in Epon. I separate the coverslip from the plastic with liquid nitrogen. The bizarre thing is that this protocol has worked well for the past few years but recently has given poor results. I'll try some of the suggestions but I'm wondering if the en bloc UA needs to be done overnight since I'm dealing with monolayers.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 16:22:48 2005



From: kevin.selkregg-at-us.vesuvius.com (by way of MicroscopyListserver)
Date: Wed, 1 Jun 2005 16:22:05 -0500
Subject: [Microscopy] viaWWW: Sodium Aluminate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (kevin.selkregg-at-us.vesuvius.com) from http://www.microscopy.org/MicroscopyListserver/MLFormMail.html on Wednesday, June 1, 2005 at 13:30:05
---------------------------------------------------------------------------

Email: kevin.selkregg-at-us.vesuvius.com
Name: Kevin Selkregg

Title-Subject: [Microscopy] [Filtered] thermal expansion coefficient

Question: Even though this question does not appear microscopy related; the thermal expansion in this situation definitely affects the microstructure of the parent material I have been studying.

Anyone have any thermal expansion coefficient data for Sodium Aluminate. I am looking specifically for the coefficient values for NaAlO2 (1 Na2O : 1 Al2O3).

Thank you.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 16:34:41 2005



From: Nestor J. Zaluzec :      zaluzec-at-microscopy.com
Date: Wed, 1 Jun 2005 16:33:59 -0500
Subject: [Microscopy] NIST-MAS-AMAS VPSEM-ESEM Workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The Following Workshop Meeting Announcement which was posted on the MAS Listserver may also be of interest
to the Microscopy Listserver readers.

Nestor.
Your Friendly Neighborhood SysOp

*****************************************************************************

NIST-MAS-AMAS Workshop on Variable Pressure and Environmental Scanning
Electron Microscopy (VPSEM-ESEM) for Imaging and Microanalysis: Roadmap 3
{http://www.nist.gov/esem}

Place: Lecture Room A and Poster Hall
National Institute of Standards and Technology,
100 Bureau Drive, Gaithersburg, Maryland 20899
Dates: November 2 4, 2005

Co-organizers:
NIST Surface and Microanalysis Science Division (837)
NIST Precision Engineering Division (821)
Microbeam Analysis Society (MAS) {http://www.microbeamanalysis.org/}
Australian Microbeam Analysis Society (AMAS) {http://www.microscopy.org.au/}

From November 2 to 4, 2005, NIST will co-host a workshop on Variable
Pressure Scanning Electron Microscopy (VPSEM) and Environmental SEM (ESEM)
in conjunction with the Microbeam Analysis Society (MAS) and the Australian
Microbeam Analysis Society (AMAS). Building on the highly successful AMAS
Roadmap 1 and 2 workshops on VPSEM-ESEM organized by Brendan Griffin, this
VPSEM-ESEM workshop, which constitutes Roadmap 3, will provide an
opportunity for technical presentations (platform and poster),
demonstrations, and discussions on leading topics in microscopy and
microanalysis in VPSEM-ESEM. Special attention will be given to new
breakthrough areas, such as combined VPSEM-ESEM-FIB (focused ion beam)
instrumentation. Discussions will attempt to identify those critical
issues where research is needed for further progress in VPSEM-ESEM.
The program as it has been currently developed is attached below. We are
seeking interested participants, proposed titles for presentations, as well
as suggestions for additional topics and invited keynote speakers to
include. In keeping with past NIST-MAS workshops, there will be no cost
for attendance, but pre-registration is required so that the available
lecture room space is not overwhelmed and NIST security requirements can be
satisfied.

NIST-MAS-AMAS VPSEM-ESEM Workshop
Tentative Program
Wednesday, Nov 2, 2005

AM: Electron and ion interactions in the gas environment
Organizer: Brad Thiel (SUNY-Albany)
Topics: fundamentals, charge control, radiation damage

PM: Contrast mechanisms in VPSEM-ESEM
Organizer: David Joy (Univ. Tennessee) and Brendan Griffin (U. Western
Australia)

Thursday, Nov. 3, 2005
AM: VPSEM-ESEM breakthrough areas:
1. Critical dimension VPSEM-ESEM microscopy
Organizer: Mike Postek/Andras Vladar (NIST)
2. VPSEM-ESEM FIB
Organizer/Invited Speaker: Lucille Giannuzzi

PM: X-ray microanalysis in the VPSEM-ESEM
Organizer: Dale Newbury (NIST)
Invited speaker: Eric Doehne (Getty Conservation Laboratory)

PM: Friday, Nov. 4, 2005
AM: Dynamic experiments in VPSEM-ESEM
Organizer: Scott Wight (NIST)

PM: Roadmap 3: Critical issues in VPSEM-ESEM
Organizer: Brendan Griffin (U. Western Australia)

Please contact dale.newbury-at-nist.gov


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 16:50:06 2005



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Wed, 01 Jun 2005 17:49:03 -0400
Subject: [Microscopy] Re: Re: via WWW: osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Pat,

Funny you should mention fixation with a mixture of
glutaraldehyde and paraforrmaldehyde. I just came
out of the scope where I looked at some samples of
drosophila egg chambers fixed with such a mixture,
followed by overnight incubation in uranyl acetate
(0.5%). Membrane contrast was unbelievable, but
also the preservation of cytoplasmic elements such
as ribosomes, actin filaments and clathrin coats. Why
don't we use that technique for all our EM preps?!

Marc

On Wednesday, June 1, 2005, at 12:06 PM, Pat Connelly wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Greetings,
} } It is really something to see one's name in print, especially
} } for being the person to go to for information on a
} } particular technique! Thanks Paul.
} }
} } With tissue cultured cells I have found it critical to fix them as
} } soon as possible after they come out of the incubator.
} } DO NOT WASH them! Simply decant the growth
} } medium then immediately and gently flood the cells with
} } the fixative.
} }
} } The fixative that works well on many different types of cultured
} } cells as well as pelleted material and small tissue pieces is as
} } follows:
} } 1% Glutaraldehyde+1% Osmium tetroxide+0.05 M Phosphate Buffer
} } -at-6.2 pH on ice and kept dark. Usually 45 minutes is fine.
} }
} } Yes, this fixative will start to oxidize. To avoid this have all the
} } components
} } on ice; mix in the glutaraldehyde immediately before the fixative is
} } needed
} } and allow sufficient volume to fix both the cells and the small
} } amount of
} } protein that is left on the cells after decanting the medium.
} }
} } Dr. Tilney INSISTS that the Electron Microscopy Sciences
} } 8% Glutaraldehyde in the 10 ml vials be used
} } at all times for his actin studies and that it be kept in
} } a scintillation vial and discarded one week after opening
} } the vial. The only reason for using the EMS product is that
} } we always get excellent results and he refuses to even try
} } another company.
} }
} } With the phosphate buffer one needs to wash well with cold
} } (best available) water at least 3 times over 20 - 30 min. time
} } AND remember to rinse the entire vessel (top too)
} } at least once before adding cold 1% UA in water
} } overnight in the refrigerator to avoid the dreaded uranyl-phosphate
} } crystals.There is no light in our refrigerator so I do not need to
} } worry
} } if it really goes out when the door is closed.
} }
} } If the cells are to be fixed in the flask/petri dish for face-on
} } sections,
} } an ethanol dehydration is used and Ladd's LX-112 as an epon
} } substitute. These do not melt the plastic.
} } All other cases or cells grown in "Pernanox" dishes
} } are acetone dehydrated and any epon substitute can be used.
} }
} } Contact me off-line if more information is needed.
} } If there are several similar questions I'll post them.
} }
} } If this fixation does not show what you wish with the membranes
} } try using an objective apperature that is a size smaller than is
} } usually used in the TEM.
} }
} } Pat Connelly psconnel-at-sas.upenn.edu
} } Dept. of Biology, University of Pennsylvania
} } Philadelphia, PA 19104-6018
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 16:53:58 2005



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Wed, 01 Jun 2005 16:53:14 -0500
Subject: [Microscopy] Re: gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

gee elaine, my first reaction was that this is an interesting question,
and that i'm sure that speculation as to this new instrumentation will
be running rife over your question. after checking the webpage you
referred to, i would even say the speculation will be royal....

i couldn't help that - please, don't ask me to apologize.... the devil
made me do it, ok....

paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-9354
Cell:204-781-1502
Fax:204-789-3926



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 17:11:51 2005



From: JoAnn Buchanan :      redhair-at-stanford.edu
Date: Wed, 01 Jun 2005 15:11:07 -0700
Subject: [Microscopy] osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Robert. You have gotten some good suggestions so far. I might add
a few things:
You should try the reduced osmium- potassium ferricyanide with cacodylate
buffer instead of phosphate and you can get rid of the ppt problem.
Leave your cells in serum until you are ready to fix and keep them warm
until ready to fix(right out of incubator). Sick cells will never look good
no matter what you do.
I embed cultured hippocampal neurons in Chang embedding molds then dissolve
away the whole glass coverslip after embedding. Dissolve the coverslip with
hydrofluoric acid(under hood)- that way you have the whole coverslip and
lots of cells. If the coverslips are coated with matrigel they are harder
to work with the nitrogen method.

For the thin sections, I stain in 5% aqueous UA a for 20 minutes or longer.
Follow with a shorter(1-2 mins) in fresh lead stain. I do not do a
prolonged UA stain en bloc-I microwave everything.
Use a lower kV or smaller aperture to image cells.
I would be happy to send you a detailed protocol.
Good luck, JoAnn

Question: Thanks for the responses but I guess I did leave out some
details. I'm processing monolayers grown on glass coverslips. I do use en
bloc staining with uranyl acetate. I'm using 2% aqueous UA for 45-60
minutes at room temperature. I use similar times for the glutaraldehyde and
osmium fixations. I dehydrate with ethanol and embed in Epon. I separate
the coverslip from the plastic with liquid nitrogen. The bizarre thing is
that this protocol has worked well for the past few years but recently has
given poor results. I'll try some of the suggestions but I'm wondering if
the en bloc UA needs to be done overnight since I'm dealing with monolayers.



Department of Molecular and Cellular Physiology
Stanford University School of Medicine
Stanford, CA 94305
650-723-5856



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 17:19:37 2005



From: Webster, Paul :      PWebster-at-hei.org
Date: Wed, 01 Jun 2005 15:16:16 -0700
Subject: [Microscopy] Re: osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Is reduced osmium prepared with ferrocyanide or ferricyanide? I know which
one I use.

Regards,

Paul Webster.




On 6/1/05 3:11 PM, "JoAnn Buchanan" {redhair-at-stanford.edu} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Dear Robert. You have gotten some good suggestions so far. I might add
} a few things:
} You should try the reduced osmium- potassium ferricyanide with cacodylate
} buffer instead of phosphate and you can get rid of the ppt problem.
} Leave your cells in serum until you are ready to fix and keep them warm
} until ready to fix(right out of incubator). Sick cells will never look good
} no matter what you do.
} I embed cultured hippocampal neurons in Chang embedding molds then dissolve
} away the whole glass coverslip after embedding. Dissolve the coverslip with
} hydrofluoric acid(under hood)- that way you have the whole coverslip and
} lots of cells. If the coverslips are coated with matrigel they are harder
} to work with the nitrogen method.
}
} For the thin sections, I stain in 5% aqueous UA a for 20 minutes or longer.
} Follow with a shorter(1-2 mins) in fresh lead stain. I do not do a
} prolonged UA stain en bloc-I microwave everything.
} Use a lower kV or smaller aperture to image cells.
} I would be happy to send you a detailed protocol.
} Good luck, JoAnn
}
} Question: Thanks for the responses but I guess I did leave out some
} details. I'm processing monolayers grown on glass coverslips. I do use en
} bloc staining with uranyl acetate. I'm using 2% aqueous UA for 45-60
} minutes at room temperature. I use similar times for the glutaraldehyde and
} osmium fixations. I dehydrate with ethanol and embed in Epon. I separate
} the coverslip from the plastic with liquid nitrogen. The bizarre thing is
} that this protocol has worked well for the past few years but recently has
} given poor results. I'll try some of the suggestions but I'm wondering if
} the en bloc UA needs to be done overnight since I'm dealing with monolayers.
}
}
}
} Department of Molecular and Cellular Physiology
} Stanford University School of Medicine
} Stanford, CA 94305
} 650-723-5856
}
}



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 17:27:10 2005



From: Tamara Howard :      thoward-at-unm.edu
Date: Wed, 1 Jun 2005 16:26:28 -0600 (MDT)
Subject: [Microscopy] Re: gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Odd that there aren't any images associated with the websites lauding
Naessens, don't you think? At least, not on any of the first page of
Google hits I came up with.

Tamara

On Wed, 1 Jun 2005, Elaine Humphrey wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Hello everyone
} I have just had a request for the use of a somatoscope developed by Gaston
} Naessens as they want colour at 30,000 times.
}
} I was wondering if someone was pulling my leg but when I did a google search
} I got
} http://www.sumeria.net/tech/naessens.html
} and I am still wondering if someone has a huge hoax going. Has anyone got
} experience with this type of microscope?
}
} Elaine
}
}
} --
} Dr. Elaine Humphrey
} Director, BioImaging Facility
} President, Microscopy Society of Canada (2003-2005)
} University of British Columbia
} 6270 University Blvd, mail-stop Botany
} Vancouver, BC
} CANADA, V6T 1Z4
} Phone: 604-822-3354
} FAX: 604-822-6089
} e-mail: ech-at-interchange.ubc.ca
} website: www.emlab.ubc.ca
}
}

|--------------------------------------------------|
Tamara Howard
Department of Cell Biology and Physiology
University of New Mexico - Health Sciences Center
Albuquerque, NM 87131
thoward-at-unm.edu
|--------------------------------------------------|


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 17:32:28 2005



From: Christopher Gilpin :      christopher.gilpin-at-utsouthwestern.edu
Date: Wed, 1 Jun 2005 17:31:43 -0500
Subject: [Microscopy] Re: gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Elaine,
To save another exile from speculation ( Royal or otherwise) I look forward
to a report on the demo when you orgainse it. :)

Chris



Christopher J Gilpin Ph.D.
Assistant Professor
Director, Molecular and Cellular Imaging Facility
K1.A04
UT Southwestern Medical Center at Dallas
5323 Harry Hines Boulevard
Dallas, TX 75390-9039
Phone +1 214 648 2827
Fax +1 214 648 6408

-----Original Message-----
} From: paul r hazelton [mailto:paul_hazelton-at-umanitoba.ca]
Sent: Wednesday, June 01, 2005 4:53 PM
To: Elaine Humphrey
Cc: microscopy-at-msa.microscopy.com

gee elaine, my first reaction was that this is an interesting question, and
that i'm sure that speculation as to this new instrumentation will be
running rife over your question. after checking the webpage you referred
to, i would even say the speculation will be royal....

i couldn't help that - please, don't ask me to apologize.... the devil made
me do it, ok....

paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-9354
Cell:204-781-1502
Fax:204-789-3926




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 1 21:20:15 2005



From: Rosemary White :      Rosemary.White-at-csiro.au
Date: Thu, 02 Jun 2005 12:20:31 +1000
Subject: [Microscopy] Re: Re: gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hmm, the only new optics I've heard about recently, also from Canada, is to
do with the liquid crystal thin lenses developed by Galstian and Presnyakov

Vladimir V. Presnyakov, Karen E. Asatryan, and Tigran Galstian (2002)
Polymer-stabilized liquid crystal for tunable microlens applications.
Optics Express 10: 865 - 870

Vladimir V. Presnyakov and Tigran V. Galstian (2005) Electrically tunable
polymer stabilized liquid-crystal lens. J. Appl. Phys. 97: 103101


} From: Tamara Howard {thoward-at-unm.edu}
} Date: Wed, 1 Jun 2005 16:26:28 -0600 (MDT)
} To: Elaine Humphrey {ech-at-interchange.ubc.ca}
} Cc: microscopy-at-msa.microscopy.com
} Subject: [Microscopy] Re: gaston naessens and somatid biology
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Odd that there aren't any images associated with the websites lauding
} Naessens, don't you think? At least, not on any of the first page of
} Google hits I came up with.
}
} Tamara
}
} On Wed, 1 Jun 2005, Elaine Humphrey wrote:
}
} }
} }
} }
-----------------------------------------------------------------------------} }
-
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------------
} } --
} }
} } Hello everyone
} } I have just had a request for the use of a somatoscope developed by Gaston
} } Naessens as they want colour at 30,000 times.
} }
} } I was wondering if someone was pulling my leg but when I did a google search
} } I got
} } http://www.sumeria.net/tech/naessens.html
} } and I am still wondering if someone has a huge hoax going. Has anyone got
} } experience with this type of microscope?
} }
} } Elaine
} }
} }
} } --
} } Dr. Elaine Humphrey
} } Director, BioImaging Facility
} } President, Microscopy Society of Canada (2003-2005)
} } University of British Columbia
} } 6270 University Blvd, mail-stop Botany
} } Vancouver, BC
} } CANADA, V6T 1Z4
} } Phone: 604-822-3354
} } FAX: 604-822-6089
} } e-mail: ech-at-interchange.ubc.ca
} } website: www.emlab.ubc.ca
} }
} }
}
} |--------------------------------------------------|
} Tamara Howard
} Department of Cell Biology and Physiology
} University of New Mexico - Health Sciences Center
} Albuquerque, NM 87131
} thoward-at-unm.edu
} |--------------------------------------------------|
}


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 04:14:47 2005



From: Allen R. Sampson :      ars-at-sem.com
Date: Thu, 2 Jun 2005 04:13:37 -0500
Subject: [Microscopy] RE: gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

A total load of crap (forgive my bluntness). Anyone wishing to argue the
point is welcome to do so directly with me, off the listserver (don't reply
to all, just me).

Allen R. Sampson, Owner
Advanced Research Systems
317 North 4th. Street
Saint Charles, Illinois 60174
phone (630) 513-7093 fax (630) 513-7092
email mailto:ars-at-sem.com web www.sem.com


On Wednesday, June 01, 2005 3:16 PM, Elaine Humphrey
[SMTP:ech-at-interchange.ubc.ca] wrote:
}
}
}
------------------------------------------------------------------------
------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
-------
}
} Hello everyone
} I have just had a request for the use of a somatoscope developed by
} Gaston Naessens as they want colour at 30,000 times.
}
} I was wondering if someone was pulling my leg but when I did a google
} search I got
} http://www.sumeria.net/tech/naessens.html
} and I am still wondering if someone has a huge hoax going. Has anyone
} got experience with this type of microscope?
}
} Elaine
}
}
} --
} Dr. Elaine Humphrey
} Director, BioImaging Facility
} President, Microscopy Society of Canada (2003-2005)
} University of British Columbia
} 6270 University Blvd, mail-stop Botany
} Vancouver, BC
} CANADA, V6T 1Z4
} Phone: 604-822-3354
} FAX: 604-822-6089
} e-mail: ech-at-interchange.ubc.ca
} website: www.emlab.ubc.ca


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 10:56:12 2005



From: Bernie Kestel :      kestel-at-anl.gov
Date: Thu, 02 Jun 2005 10:55:25 -0500
Subject: [Microscopy] High Temperature Bore Scope and Telescope:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


A co-worker asked me to assist him with the following problem.
He has metal 3 m.m. diameter discs that are about 0.25 m.m. thick. They
are radioactive due to being neutron irradiated. A fixture has been made
which resembles a punch and die to mechanically deform the discs at 300
degrees centigrade by hydralic pressure applied to a central rod about
1 m.m. In diameter. The "viewing hole" below the disc will be 1.5 m.m.
in diameter by 3 centimeters long or "deep."
What is needed is a telescopic device which can both illuminate the
surface of the disc and transmit an image of it to a viewing
device or camera which is several inches away.

Bernie Kestel
Material Science Division
Argonne National Laboratory
9700 South Cass Avenue
Argonne, Illinois, 60439



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 11:02:09 2005



From: Lesley Weston :      leswes-at-shaw.ca
Date: Thu, 02 Jun 2005 09:01:12 -0700
Subject: [Microscopy] Re: gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

It's a lovely idea - the cure for cancer and everything else - and if you
read it fast enough it almost makes sense. And it comes complete with a
conspiracy theory: "Unfortunately, Rife's success attracted the
attention and wrath of the American Medical Association (AMA) and the
powerful pharmaceutical companies - the organised opposition of the medical
fields", which makes it even more authentic.

--
Lesley Weston


} From: Elaine Humphrey {ech-at-interchange.ubc.ca}
} Date: Wed, 01 Jun 2005 13:16:14 -0700
} To: microscopy-at-msa.microscopy.com
} Subject: [Microscopy] gaston naessens and somatid biology
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
----------------------------------------------------------------------------
--} -
}
} Hello everyone
} I have just had a request for the use of a somatoscope developed by
} Gaston Naessens as they want colour at 30,000 times.
}
} I was wondering if someone was pulling my leg but when I did a google
} search I got
} http://www.sumeria.net/tech/naessens.html
} and I am still wondering if someone has a huge hoax going. Has anyone
} got experience with this type of microscope?
}
} Elaine
}
}
} --
} Dr. Elaine Humphrey
} Director, BioImaging Facility
} President, Microscopy Society of Canada (2003-2005)
} University of British Columbia
} 6270 University Blvd, mail-stop Botany
} Vancouver, BC
} CANADA, V6T 1Z4
} Phone: 604-822-3354
} FAX: 604-822-6089
} e-mail: ech-at-interchange.ubc.ca
} website: www.emlab.ubc.ca
}



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 12:10:46 2005



From: Roger Baker :      rcbaker-at-eden.infohwy.com
Date: Thu, 2 Jun 2005 12:09:47 -0500
Subject: [Microscopy] Re: High Temperature Bore Scope and Telescope:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

All you likely need is a consumer digital camera that has a video
output for a TV monitor and an small achromatic lens of the right
focal length mounted in front, and most of its surface apertured out
to shield stray light. Some Coolpix cameras are nice because the
front lens surface stays fixed while zooming or focusing.

If you mount a piece of a clean cover slip at the right angle just
outside the hole and then have a bright collimated light beam
reflected into the hole by the tilted glass, there should be little
scattering and your camera setup should get a good hole bottom
picture through the cover glass.

I am sure there are also many more fancy and expensive ways to do the
same job. -- Roger



On Jun 2, 2005, at 10:55 AM, Bernie Kestel wrote:

}
}
} ----------------------------------------------------------------------
} --------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/
} MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} ----------------------------------------------------------------------
} ---------
}
}
} A co-worker asked me to assist him with the following problem.
} He has metal 3 m.m. diameter discs that are about 0.25 m.m.
} thick. They
} are radioactive due to being neutron irradiated. A fixture has
} been made
} which resembles a punch and die to mechanically deform the discs
} at 300
} degrees centigrade by hydralic pressure applied to a central rod
} about
} 1 m.m. In diameter. The "viewing hole" below the disc will be
} 1.5 m.m.
} in diameter by 3 centimeters long or "deep."
} What is needed is a telescopic device which can both
} illuminate the
} surface of the disc and transmit an image of it to a viewing
} device or camera which is several inches away.
}
} Bernie Kestel
} Material Science Division
} Argonne National Laboratory
} 9700 South Cass Avenue
} Argonne, Illinois, 60439
}
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 14:39:47 2005



From: Gazda, Jerzy :      Jerzy.Gazda-at-ceriumlabs.com
Date: Thu, 2 Jun 2005 14:38:58 -0500
Subject: [Microscopy] Announcement - SEM/TEM Semiconductor and Nanotechnology Sample

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

SEM/TEM Semiconductor and Nanotechnology Sample Prep Workshop
June 7, 2005
Austin, TX

Cerium Laboratories LLC. and Sagitta Inc. will be holding a FREE
workshop to present applications of automated polishing to sample
preparations from semiconductor and nanotechnology materials for SEM,
TEM, and AFM based evaluations.

Two sessions will be held. The morning session will concentrate on SEM
prep, while the afternoon session will be addressing TEM and SRP-AFM
techniques. There will be short presentations and live demonstration of
automated polishing followed by SEM or TEM imaging of these samples
using equipment at CeriumLabs.
A possibility exists for preparation of specimens brought by
participants and imaging sessions on June 8 and 9. Please contact us to
set them up.

For more information or to register, please contact:
Jerzy Gazda - Jerzy.Gazda-at-ceriumlabs.com - 800-538-8450, Ext. 51453; or
Efrat Raz - Efrat.Raz-at-sagitta.com - 408-249-9060.

Please RSVP

Jerzy

******************************************************
Jerzy Gazda, Ph.D. CeriumLaboratories L.L.C.

Supervising Engineer 5204 E. Ben White
Blvd. - MS 512
Austin, TX 78741
TEL: 1-800-538-8450, Ext. 51453
jerzy.gazda-at-ceriumlabs.com
******************************************************





From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 20:12:34 2005



From: brian.leonard-at-sunhealth.org (by way of MicroscopyListserver)
Date: Thu, 2 Jun 2005 20:11:51 -0500
Subject: [Microscopy] viaWWW: Human leukocyte embedment and thin sectioning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (brian.leonard-at-sunhealth.org) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Thursday, June 2, 2005 at 16:33:18
---------------------------------------------------------------------------

Email: brian.leonard-at-sunhealth.org
Name: Brian Leonard

Organization: Sun Health Research Institute

Title-Subject: [Microscopy] [Filtered] Human leukocyte embedment and thin sectioning

Question: Does anyone have a protocol for em processing of buffy coat from human whole blood? I plan on doing post-embedding immuno on thin sections of leukocytes. The ultrastructure of the cells processed with my current method is terrible. I fractionate EDTA-treated whole blood over a Ficoll gradient to get leukocytes, and then fix the buffy coat cells at 4 deg C in 2.5% glut/4% para in 0.1M PB (or pbs). After pelleting, I resuspend the cells in low-melting agarose (37-40 deg) to put the cells in a stable matrix for the remainder of the processing. After dissecting the gel matrix into ~1mm cubes, the remainder of the processing is standard em processing for tissue sections, with final embeddment in L.R. White. Any help is much appreciated. Thanks

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 20:12:53 2005



From: lcgould-at-med.cornell.edu (by way of MicroscopyListserver)
Date: Thu, 2 Jun 2005 20:12:10 -0500
Subject: [Microscopy] viaWWW: osmium fixation of cultured cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (lcgould-at-med.cornell.edu) from http://microscopy.com/MLFormMail.html on Thursday, June 2, 2005 at 19:47:23
---------------------------------------------------------------------------

Email: lcgould-at-med.cornell.edu
Name: Leona Cohen-Gould

Organization: Joan & Sanford I. Weill Medical College

Title-Subject: [Microscopy] [Filtered] viaWWW: osmium fixation of cultured cells

Question: Robert,
For membrane preservation I use a fixation protocol given to me years ago by someone who studied photoreceptors (tons of membrane). Her recipes were as follows:
Primary fix: 2.5% glut, 4% pfa and 0.2% picric acid in 0.1M cacodylate buffer, pH 7.3
Post fix: 1% osmium tetrox., 1.5% Pot. ferricyanide (aqueous)

I also en bloc stain with Ur Ac.

for the picric acid (my office of environmental health and safety just loves me) I keep the smallest jar available with the crystals fully covered with water. I use the resultant saturated solution in my primary fix (2 ml picric acid sol'n to 40 ml fix).

The primary fix is based on one published by Somogyi and Takagi in Neuroscience Vol 7 No 7 pp1779-1783, 1982

Lee

--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 2 22:41:53 2005



From: h.cao-at-genesis.co.nz (by way of Ask-A-Microscopist)
Date: Thu, 2 Jun 2005 22:41:10 -0500
Subject: [Microscopy] AskAMicroscopist :fluorescence in skin tissues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (h.cao-at-genesis.co.nz) from http://www.microscopy.org/Ask-A-Microscopist/Ask-A-Microscopist.html on Thursday, June 2, 2005 at 22:12:27
---------------------------------------------------------------------------

Email: h.cao-at-genesis.co.nz
Name: Helen Cao

Organization: Genesis R & D Ltd

Education: Graduate College

Location: Auckland, New Zealand

Question: Dear all,

I am having some problems in examining fluorescence in skin tissues. I injected (intradermally ) a fluorescence marker into mouse skin, then fixed skin tissues in 4% PFA at 4C overnight, washed in PBS (3 times, 20min each time at 4C), 20% sucrose at 4C overnight. The skin tissues were then embedded in OCT and frozen over liquid nitrogen. No good sections could be obtained, the microanatomical structures in the sections were fragmented. I tried to freeze fresh skin (unfixed) and obtained very good morphology. Any suggestions on how to preserve the morphology and fluorescence in skin tissues? Can I freeze fresh skin tissues to get good sections, and then fix the tissues before examine the fluorescence? Or, can I fix skin tissues in 4% PFA, get paraffin sections and then examine fluorescence? Do these methods quench the fluorescence?

Thanks for your help in advance


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 08:13:15 2005



From: adriana-at-cab.cnea.gov.ar (by way of MicroscopyListserver)
Date: Fri, 3 Jun 2005 08:12:32 -0500
Subject: [Microscopy] viaWWW: TEM EDS for Philips CM200 UT

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (adriana-at-cab.cnea.gov.ar) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday, June 3, 2005 at 07:29:20
---------------------------------------------------------------------------

Email: adriana-at-cab.cnea.gov.ar
Name: Adriana CondÛ

Organization: CONICET, Argentina

Title-Subject: [Microscopy] [Filtered] MListserver: TEM EDS for Philips CM200 UT

Question:
Dear microscopists,

We are interested in buying an EDS detector for a Philips CM200 TEM with an ultratwin lens.
We would like to have user experience on the performance of the following EDS systems:

a) Noran System SIX model provided by Thermo Electron, or similar models, that have no shutter, are non-retractable but have automatic FET protection.

b) EDAX Genesis 2000 model, that is non-retractable and has no shutter protection (because of the ultratwin lens).

We are particularly concerned with the damage to the detector when the microscope operates at low magnification. We would like to know whether the detector can be permanently damaged if it receives a high dose of radiation, or if the only consequences are practical (wait until the system decays).

All comments will be appreciated.

Thank you for your help.

Adriana


----------------------------------------------------------
Adriana CondÛ
CONICET Researcher
Metals Physics Group
Centro AtÛmico Bariloche
8400 San Carlos de Bariloche
ARGENTINA
e-mail: adriana-at-cab.cnea.gov.ar
http://www.cab.cnea.gov.ar/cab/invbasica/metales
fax: +54 2944 445299
TE: +54 2944 445290
----------------------------------------------------------


---------------------------------------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 08:42:59 2005



From: Sherwood, Margaret :      MSHERWOOD-at-PARTNERS.ORG
Date: Fri, 3 Jun 2005 09:42:13 -0400
Subject: [Microscopy] AskAMicroscopist :fluorescence in skin tissues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I work in a Dermatology/Laser Research Lab and we cut frozen sections of skin
all the time for fluorescent studies and confocal work.

Peggy Sherwood
Lab Associate, Photopathology
Wellman Center for Photomedicine (W224)
Massachusetts General Hospital
55 Fruit Street
Boston, MA 02114
617-724-4839 (voice mail)
617-726-6983 (lab)
617-726-3192 (fax)
msherwood-at-partners.org



-----Original Message-----
} From: by way of Ask-A-Microscopist [mailto:h.cao-at-genesis.co.nz]
Sent: Thursday, June 02, 2005 11:41 PM
To: microscopy-at-microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by
(h.cao-at-genesis.co.nz) from
http://www.microscopy.org/Ask-A-Microscopist/Ask-A-Microscopist.html on
Thursday, June 2, 2005 at 22:12:27
---------------------------------------------------------------------------

Email: h.cao-at-genesis.co.nz
Name: Helen Cao

Organization: Genesis R & D Ltd

Education: Graduate College

Location: Auckland, New Zealand

Question: Dear all,

I am having some problems in examining fluorescence in skin tissues. I injected
(intradermally ) a fluorescence marker into mouse skin, then fixed skin tissues
in 4% PFA at 4C overnight, washed in PBS (3 times, 20min each time at 4C), 20%
sucrose at 4C overnight. The skin tissues were then embedded in OCT and frozen
over liquid nitrogen. No good sections could be obtained, the microanatomical
structures in the sections were fragmented. I tried to freeze fresh skin
(unfixed) and obtained very good morphology. Any suggestions on how to preserve
the morphology and fluorescence in skin tissues? Can I freeze fresh skin tissues
to get good sections, and then fix the tissues before examine the fluorescence?
Or, can I fix skin tissues in 4% PFA, get paraffin sections and then examine
fluorescence? Do these methods quench the fluorescence?

Thanks for your help in advance


---------------------------------------------------------------------------




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 10:11:05 2005



From: frank.karl-at-degussa.com
Date: Fri, 3 Jun 2005 11:30:58 -0400
Subject: [Microscopy] Photoshop 7 vs Photoshop CS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I'm looking for some advice. I understand Adobe has issued Photoshop 8 and
called it Photoshop CS2. I am currently using Photoshop 6 with Reindeer
Graphics's Fovea Pro V3. Is anyone actually using Photoshop CS2 and is it
better than Photoshop 7 and will my Povea Pro 3 work with it. I don't
want to be too far behind the curve as technology changes, but I don't want
to be the public debuger either!

Thanks!!

Frank Karl
Degussa Corporation
Akron Technical Center
3500 Embassy Parkway
Suite 100
Akron, Ohio 44333


330-668-2235 Ext. 238


This e-mail transmission, and any documents, files or previous e-mail
messages attached to it may contain information that is confidential or
legally privileged. If you are not the intended recipient, or a person
responsible for delivering it to the intended recipient, you are hereby
notified that you must not read this transmission and that any disclosure,
copying, printing, distribution or use of any of the information contained
in or attached to this transmission is STRICTLY PROHIBITED. If you have
received this transmission in error, please immediately notify the sender
by telephone or return e-mail and delete the original transmission and its
attachments without reading or saving in any manner. Thank you.



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 10:43:44 2005



From: Pat Connelly :      psconnel-at-sas.upenn.edu
Date: Fri, 3 Jun 2005 11:25:44 -0400
Subject: [Microscopy] Re: via WWW: Human leukocyte embedding and thin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Brian:
I have fixed leukocytes that were grown on coverslips but I think that the
same technique may solve your problem.
First: keep PBS away from any cells that you ever wish to do EM on.
This makes a noticeable difference.
LM /confocal investigators do not seem to worry about it.
Second: try fixing the buffy coat at room temperature unless it is already
cold - then cool it down if protocol requires it.
Third: the fixative may work better if it was less concentrated.
1%glutaraldehyde seems to be the top concentration for immuno.
studies but freshly prepared (para)formaldehyde is used in many
labs by itself or in combination with glutaraldehyde.
Others will be sure to remark on this point.
You are welcome,
Pat Connelly
Dept. of Biology
Univ. of Pennsylvania
Philadelphia, PA 19104-6018
psconnel-at-sas.upenn.edu
===================
Below is the result of your feedback form (NJZFM-ultra-55).

Email: brian.leonard-at-sunhealth.org
Name: Brian Leonard
Organization: Sun Health Research Institute
Title-Subject: [Microscopy] Human leukocyte embedment and thin sectioning
Question: Does anyone have a protocol for em processing of buffy coat from
human whole blood? I plan on doing post-embedding immuno on thin sections
of leukocytes. The ultrastructure of the cells processed with my
current method is terrible. I fractionate EDTA-treated whole blood
over a Ficoll gradient to get leukocytes, and then fix the buffy coat
cells at 4 deg C in 2.5% glut/4% para in
0.1M PB (or pbs). After pelleting, I resuspend the cells in low-melting agarose
37-40 deg.) to put the cells in a stable matrix for the remainder of
the processing.After dissecting the gel matrix into ~1mm cubes, the
remainder of
the processing is standard EM processing for tissue sections, with final
embeddment in L.R. White.
Any help is much appreciated. Thanks


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 10:52:07 2005



From: David Elliott :      Elliott-at-arizona.edu
Date: Fri, 3 Jun 2005 08:51:20 -0700
Subject: [Microscopy] Re: Photoshop 7 vs Photoshop CS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Photoshop 8 came out with CS1. I am using it. I use lots of complex
scripts and the Photoshop 8 scripting was worth my money.
It has lots of other great things, but most of them are of no use to
microscopists (unless you want to forge your data :-)
David


On Jun 3, 2005, at 8:30 AM, frank.karl-at-degussa.com wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} I'm looking for some advice. I understand Adobe has issued Photoshop
} 8 and
} called it Photoshop CS2. I am currently using Photoshop 6 with
} Reindeer
} Graphics's Fovea Pro V3. Is anyone actually using Photoshop CS2 and
} is it
} better than Photoshop 7 and will my Povea Pro 3 work with it. I don't
} want to be too far behind the curve as technology changes, but I don't
} want
} to be the public debuger either!
}
} Thanks!!
}
} Frank Karl
} Degussa Corporation
} Akron Technical Center
} 3500 Embassy Parkway
} Suite 100
} Akron, Ohio 44333
}
}
} 330-668-2235 Ext. 238
}
}
} This e-mail transmission, and any documents, files or previous e-mail
} messages attached to it may contain information that is confidential or
} legally privileged. If you are not the intended recipient, or a person
} responsible for delivering it to the intended recipient, you are hereby
} notified that you must not read this transmission and that any
} disclosure,
} copying, printing, distribution or use of any of the information
} contained
} in or attached to this transmission is STRICTLY PROHIBITED. If you have
} received this transmission in error, please immediately notify the
} sender
} by telephone or return e-mail and delete the original transmission and
} its
} attachments without reading or saving in any manner. Thank you.
}
}



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 10:54:58 2005



From: Ronald Anderson :      randerson20-at-tampabay.rr.com
Date: Fri, 03 Jun 2005 11:54:07 -0400
Subject: [Microscopy] Re: Photoshop 7 vs Photoshop CS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Frank,

I'm using Adobe Creative Suite CS2 Premium to produce Microscopy Today.
The package includes Photoshop CS2. The first thing I did after
installing the suite was to install my Fovea Pro 3 filters, etc.
Everything works perfectly. What you see in Microscopy Today is
testament. The July issue, in the works, will be the first one done all
in CS2. There are some very valuable improvements in PS 7 (CS) and more
in PS 8 (CS2). Adobe is beginning to recognize the scientific market to
our considerable benefit.

I have no interest in Adobe other than being a very satisfied user.

Ron Anderson, Editor
Microscopy Today


frank.karl-at-degussa.com wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 12:23:26 2005



From: Joe Kulik :      juk12-at-psu.edu
Date: Fri, 3 Jun 2005 14:35:18 -0400
Subject: [Microscopy] TEM: Spare parts for Philips instrument

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I'm using Photoshop CS, which is Photoshop 8 (the number shown when it first
opens), with Fovea Pro 3 and Optipix 3 with no problems whatsoever. I don't
know the numerical designation for Photoshop CS2 but will learn more about
it tomorrow in a class.

Damian Neuberger, Ph.D.



We have a Philips EM420T on which one of the push-button function
selector switches (The M button for image mode) is starting to go bad.
(It sometimes does not engage smoothly and the OL current then changes.)

Replacement parts for these old instruments are depressingly expensive.

Does anyone have a decommissioned Philips TEM of the same vintage, and,
if so, would you be willing to give us (or sell us at a reasonable
price) spare parts like the aforementioned switch?

Or can anyone suggest a less expensive alternative than buying a $1500
switch from FEI?

Thanks to anyone with any suggestions!
_________________________________

Joseph Kulik
Research Associate
Materials Research Institute
The Pennsylvania State University
194 MRI Bldg
University Park, PA 16802

Telephone: 814-865-0344
Fax: 814-863-8561
e-mail: juk12-at-psu.edu
_________________________________




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 15:32:00 2005



From: Ronald Anderson :      randerson20-at-tampabay.rr.com
Date: Fri, 03 Jun 2005 16:31:13 -0400
Subject: [Microscopy] Re: Re: Photoshop 7 vs Photoshop CS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

David is right. Photoshop CS2 is PS 9, not PS 8 as I said in my previous
post. Check out Adobe site to see a list of new and improved features.

Ron Anderson


David Elliott wrote:

}
}
} ------------------------------------------------------------------------------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
}
} Photoshop 8 came out with CS1. I am using it. I use lots of complex
} scripts and the Photoshop 8 scripting was worth my money.
} It has lots of other great things, but most of them are of no use to
} microscopists (unless you want to forge your data :-)
} David
}
}
} On Jun 3, 2005, at 8:30 AM, frank.karl-at-degussa.com wrote:
}
} }
} }
} } -----------------------------------------------------------------------
} } -------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } America
} } To Subscribe/Unsubscribe --
} } http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------
} } --------
} }
} } I'm looking for some advice. I understand Adobe has issued
} } Photoshop 8 and
} } called it Photoshop CS2. I am currently using Photoshop 6 with
} } Reindeer
} } Graphics's Fovea Pro V3. Is anyone actually using Photoshop CS2 and
} } is it
} } better than Photoshop 7 and will my Povea Pro 3 work with it. I don't
} } want to be too far behind the curve as technology changes, but I
} } don't want
} } to be the public debuger either!
} }
} } Thanks!!
} }
} } Frank Karl
} } Degussa Corporation
} } Akron Technical Center
} } 3500 Embassy Parkway
} } Suite 100
} } Akron, Ohio 44333
} }
} }
} } 330-668-2235 Ext. 238
} }
} }
} } This e-mail transmission, and any documents, files or previous e-mail
} } messages attached to it may contain information that is confidential or
} } legally privileged. If you are not the intended recipient, or a person
} } responsible for delivering it to the intended recipient, you are hereby
} } notified that you must not read this transmission and that any
} } disclosure,
} } copying, printing, distribution or use of any of the information
} } contained
} } in or attached to this transmission is STRICTLY PROHIBITED. If you have
} } received this transmission in error, please immediately notify the
} } sender
} } by telephone or return e-mail and delete the original transmission
} } and its
} } attachments without reading or saving in any manner. Thank you.
} }
} }
}
}
}




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 17:03:34 2005



From: Hong Yi :      hyi-at-emory.edu
Date: Fri, 3 Jun 2005 18:02:51 -0400
Subject: [Microscopy] (Microscopy) Pre-embedding Immunogold Labeling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear All:

Could I ask those who do pre-embedding immunogold labeling routinely
to contact me off-line? I would like to discuss with you a few issues
regarding silver enhancement. Thank you in advance.

Hong
Emory SOM EM



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 3 22:26:45 2005



From: DrJohnRuss-at-aol.com
Date: Fri, 3 Jun 2005 23:25:56 EDT
Subject: [Microscopy] Re: Photoshop 7 vs Photoshop CS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


In a message dated 6/3/05 11:14:03 AM, frank.karl-at-degussa.com writes:

} I'm looking for some advice. I understand Adobe has issued Photoshop 8
} and
} called it Photoshop CS2. I am currently using Photoshop 6 with Reindeer
} Graphics's Fovea Pro V3. Is anyone actually using Photoshop CS2 and is
} it
} better than Photoshop 7 and will my Povea Pro 3 work with it. I don't
} want to be too far behind the curve as technology changes, but I don't
} want
} to be the public debuger either!


CS2 (aka Photoshop 9) has quite a few new capabilities, but whether you need
them is hard to say. The handling of 16 bit images is quite a bit smoother and
more complete than in PS7, and the layers capability is also much better. The
Fovea Pro plugins work just fine with CS2, but you will have to re-record any
actions you may have set up to automate Photoshop as they do not translate
smoothly from PS9 to PS9.

John Russ


From MicroscopyL-request-at-ns.microscopy.com Sat Jun 4 08:19:44 2005



From: dgrimm-at-tulane.edu (by way of MicroscopyListserver)
Date: Sat, 4 Jun 2005 08:19:02 -0500
Subject: [Microscopy] viaWWW: Used Microscopes Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (dgrimm-at-tulane.edu) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday, June 3, 2005 at 09:50:36
---------------------------------------------------------------------------

Email: dgrimm-at-tulane.edu
Name: Deborah Grimm

Organization: Tulane University

Title-Subject: [Microscopy] [Filtered] Used Microscopes

Question:

AMRAY 1700 SEM, has not been under vacuum for over 1 year, but should be operational, as-is where-is, any reasonable offer accepted.

JEOL JSM 820 SEM, analog, probably a parts machine. Any reasonable offer accepted.

We need to remove these microscopes from our lab by June 17th. Please e-mail me if you have any questions.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sat Jun 4 08:20:56 2005



From: DrJohnRuss-at-aol.com
Date: Sat, 4 Jun 2005 09:19:58 EDT
Subject: [Microscopy] correction for latest Microscopy Today

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The latest issue of Microscopy Today carries another of Jerry Sedgewick’s
articles about Photoshop techniques. Like every one of his others, it contains
much that is flat-out wrong or misleading, omits important information, and even
the parts that will actually work are by far not the best, easiest, or most
useful ways to accomplish the same purpose. I’ve tried in the past to send
corrective information (which has never gotten any response from either Sedgewick
or from Ron Anderson, who continues to accept these articles even though
several people besides me have complained about them and pointed out the numerous
errors). I suppose the need to fill up the magazine with unreviewed and
blatantly incorrect stuff outweighs any other criteria, which is unfortunate because
it reflects badly on other articles that are contributed by serious and
respectable scientists and doesn't help the magazine achieve respectability.

Anyway, to the subject: The article concerns methods for combining stereo
pair images to produce color anaglyphs. It is important to note some errors.
First, the drivel about converting 16 bit to 8 bit images is incorrect. You can
directly construct anaglyphs with either 8 or 16 bit images (or even a mixture,
which is admittedly unlikely), since Photoshop’s latest two versions (8 and 9,
also known as CS and CS2) handle 16 bit images just fine. And even if you do
want to convert a 16 bit image to 8 bits, the procedure described in the
article and shown in Figure 1 is wrong. That is the method (which was published
some time ago in this same magazine by Brent Neal) applies to converting a 12 bit
image to 16 bits, and has nothing whatever to do with the subject at hand. If
for some reason you do want to convert a 16 bit image to 8 bits, just select
Image-} Mode-} 8 bits per channel. Done.

But the main part of the article concerns putting together two images into
the color channels of a composite anaglyph. There is no reason whatever to use
the cumbersome procedure described, with extra black images and the merge
channels routine and awkward steps to align things. Just do the following (this
works with versions of Photoshop at least back to 6 - I haven’t tried older ones
but it probably works with 5 as well):
1) Select-} All and Edit-} Copy the left eye image to place it on the clipboard
2) Change the mode of the right eye image from Grayscale to RGB Color. This
will not change the appearance of the image on the screen, but if you look at
the Channels palette you will see that there are now three (RGB) channels all
containing the same data.
3) Click on the Red channel in the palette and paste in the image. The
Sedgewick article omits the important fact that the convention for viewing stereo
anaglyphs is to use the red channel for the left eye and either green or blue
for the right, and that is the way glasses are constructed. It’s difficult to
tell in the printed article, but it appears that the example shown in Figure 2
was put together backwards.
4) Click the eye symbol for RGB to display all of the channels. The original
right eye image is now shown in both green and blue, which works with either
the red/green or red/blue glasses and is brighter and easier to see than using
just one of those two channels.

You may be finished now, and can crop the image if desired and save it. But
if the alignment of the images is imperfect - either there is a vertical shift
between the images, or one is slightly rotated, or you just want to change the
horizontal displacement to alter the impression of depth - you can do it now
without any of the “take it apart, fiddle around, and try again†advice in
the Sedgewick article. The red channel is still selected from the steps above.
You can shift it (this is usually done more controllably with the arrow keys
than with the mouse - use the arrows to move the red channel a pixel at a time,
or depress the shift key as well to move it ten pixels at a time), or even
rotate it using the Edit-} Transform function. The key here is that you are
viewing the full color anaglyph through your glasses, because you turned on the RGB
visibility in step 4, and can therefore immediately see what your shifting of
the image does visually, which gives precise control. All of the stuff in the
Sedgewick article about locking or unlocking the layer, splitting and merging
the channels, creating an additional image, and so on, is just unnecessary and
confusing, and makes the task more difficult.

John Russ




From MicroscopyL-request-at-ns.microscopy.com Sat Jun 4 10:03:24 2005



From: km602223-at-comcast.net
Date: Sat, 04 Jun 2005 15:02:38 +0000
Subject: [Microscopy] Nikon Fluophot

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Does anyone have an instruction manual for a Nikon Fluophot microscope?
Thank you,
Kathleen


From MicroscopyL-request-at-ns.microscopy.com Sat Jun 4 15:30:24 2005



From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Sat, 4 Jun 2005 16:29:42 -0400 (EDT)
Subject: [Microscopy] Re: Re: Photoshop 7 vs Photoshop CS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

There is a demo version of CS2 that you can download and run. It has a
lot of fun new stuff, a lot definitely NOT useful for scientific imaging.
But two features for making figures that may be worth the ungrade are 1.)
the ability to shrink and reenlarge images without throwing out the
original pixels (I think this is from Illustrator) and 2.) a more sensible
font selection tool.

} Is anyone actually using Photoshop CS2 and is
} } it
} } better than Photoshop 7 and will my Povea Pro 3 work with it. I don't
} } want to be too far behind the curve as technology changes, but I don't
} } want
} } to be the public debuger either!

_________________________________________
Michael Cammer
Analytical Imaging Facility and
Dept. ASB Biophotonics Innovation Laboratory
Albert Einstein College of Medicine
1300 Morris Park Avenue, Bronx, NY 10461
718-430-2890 Fax 718-430-8996
work: http://www.aecom.yu.edu/aif/
personal: http://coxcammer.com/




From MicroscopyL-request-at-ns.microscopy.com Sat Jun 4 16:31:31 2005



From: Berns Buenaobra :      bbuenaobra-at-nip.upd.edu.ph
Date: Sun, 5 Jun 2005 05:32:39 -0700
Subject: [Microscopy] Device DDK

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello:

I'm trying to customize a control application for a BX Olympus motorized
microscope but is short of a device driver kit for it (DDK). My problems
are determining the exact communications protocol to connect the serial
port to the controlling box.

Any one has the DDK or the protocol description?

Regards,

Berns B.



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 09:27:32 2005



From: Geoff McAuliffe :      mcauliff-at-umdnj.edu
Date: Mon, 06 Jun 2005 10:26:22 -0400
Subject: [Microscopy] Re: Re: gaston naessens and somatid biology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Elaine et al:

Hoax/fraud/BS, pick one. Certainly not true.You might quote the
requestor an outrageous price and see how they respond. :-)

Geoff

} On Wed, 1 Jun 2005, Elaine Humphrey wrote:
}
} } ------------------------------------------------------------------------------
} }
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} } http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } -------------------------------------------------------------------------------
} }
} }
} } Hello everyone
} } I have just had a request for the use of a somatoscope developed by
} } Gaston Naessens as they want colour at 30,000 times.
} }
} } I was wondering if someone was pulling my leg but when I did a google
} } search I got
} } http://www.sumeria.net/tech/naessens.html
} } and I am still wondering if someone has a huge hoax going. Has anyone
} } got experience with this type of microscope?
} }
} } Elaine
} }
} }
} } --
} } Dr. Elaine Humphrey
} } Director, BioImaging Facility
} } President, Microscopy Society of Canada (2003-2005)
} } University of British Columbia
} } 6270 University Blvd, mail-stop Botany
} } Vancouver, BC
} } CANADA, V6T 1Z4
} } Phone: 604-822-3354
} } FAX: 604-822-6089
} } e-mail: ech-at-interchange.ubc.ca
} } website: www.emlab.ubc.ca
} }
} }


--
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 10:31:19 2005



From: Skage Hem :      SHem-at-laurentian.ca
Date: Mon, 06 Jun 2005 11:29:58 -0400
Subject: [Microscopy] Sample charging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello there,
First I wish to express my gratitude towards those list members who
helped me out with the electrical schematics and manuals for the
Cambridge S-120. Thanks alot!

Secondly, the samples seem to be charging/arching in a pulsating manner, even when the sample is thoroughly grounded. I know the column is somewhat dirty, do anyone know if this could be related to the trouble. If not do anyone have any ideas as what might be the cause.

Any advice is greatly appreciated.

Skage Hem





_______________________________________________________________

Skage Hem, Ph.D.
Research Scientist
CAF, Department of Earth Sciences
Laurentian University
Ramsey Lake Road
Sudbury, ON
Canada
P3E 2C6

ph. 705-562-7321
fax 705-675-4898

http://laurentian.ca/geology/Research/hem.html




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 11:03:32 2005



From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Mon, 06 Jun 2005 11:00:14 -0500
Subject: [Microscopy] Re: Sample charging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Are you certain the sample is grounded? We have some coated samples that it
seems we can never put enough gold on to eliminate charging. The samples
are insulating and rough. I can only examine small aggregates of particles
without encountering too much charging. I would try looking at a piece of
well-grounded, clean metal to be sure there is no chance of it being the
sample.

I went through that exercise with a dirty column a couple years ago. The
service engineer strongly suspected the sample, but when I loaded a bare
sample holder and still got drifting, he was finally convinced. He tore our
column down ALL the way and found some contamination deep within. I think
our scope was seven-years-old at the time. The location does not get dirty
so fast that it requires cleaning at every preventative maintenance, but we
will want to watch for problems and probably clean it at least every 4 or 5
years.

Warren

At 10:29 AM 06/06/05, you wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

-------------------------------------------
No files should be attached to this message
-------------------------------------------
Warren E. Straszheim, Ph.D.
Materials Analysis and Research Lab
Iowa State University
46 Town Engineering
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
Web: www.marl.iastate.edu

Scanning electron microscopy, x-ray analysis, and image analysis of materials
Computer applications and networking



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 11:08:47 2005



From: Pat Connelly :      psconnel-at-sas.upenn.edu
Date: Mon, 6 Jun 2005 11:51:04 -0400
Subject: [Microscopy] Sample charging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The samples seem to be charging/arching in a pulsating manner, even
when the sample is thoroughly grounded. I know the column is somewhat
dirty, do anyone know if this could be related to the trouble. If not
do anyone have any ideas as what might be the cause.
Skage Hem {SHem-at-laurentian.ca}
============
Skage,
Does your Cambridge S-120 have a spark plug or someplace where there
is a gap like a spark plug? [Stop laughing!]

I have a pulsing when a fleck of carbon gets built up inside the gap
of the spark plug in my Philips 200. It took weeks to find this
answer many years ago and it has occured only once in the last 10
years.
The cure is to run a piece of paper through the gap which is very
small to remove the carbon without needing to remove the spark plug
or readjust the gap space.

Pat Connelly psconnel-at-sas.upenn.edu
The University of Pennsylvania
Department of Biology
Philadelphia, PA 19104-6018
215-898-7145


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 11:21:44 2005



From: Ralph Common :      Ralph.Common-at-hc.msu.edu
Date: Mon, 6 Jun 2005 12:20:18 -0400
Subject: [Microscopy] Focus Extender 1.0

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am considering buying Reindeer Graphics Focus Extender 1.0, and would like to hear feedback from actual users. A search of the Web turned up little except copies of the original press release. I would like to know about ease of use, how well the program handles various types of subjects, and how it stacks up (pun intended) against similar programs. On a related topic, does anybody know of a source for a computer driven Z-drive that doesn't cost a fortune?

Ralph Common
Michigan State University
Division of Human Pathology



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 11:25:11 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 06 Jun 2005 09:24:24 -0700
Subject: [Microscopy] [MICROSCOPY] specimen heating up

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I tried two other sets of specimens and they all heated up.
The ambient temperature is 70F. The heated temperature is
95F.

One specimen is a sputter coated (Pt) #1 cover slip stuck on
a pin stub via Nisshin EMS carbon tab. The slip is offset
slightly so I can ground it using Pella colloidal silver.

When I put this in the chamber, no beam, for about 10 minutes,
I pull it out and it is hot.

Then I did three Al film on silicon pieces on stubs which were attached using
Buehler wax but grounded using the Pella colloidal silver.
This large holder also got up to 95F in chamber vacuum without
any beam.

After some amount of time in air, the specimens seem to not
heat up. What is going on?

Strange. But it messes up image collection due to large
amount of drift.

gary g.



At 08:30 AM 5/31/2005, you wrote:
} Just a wild guess, but would the sample have heated up the same if mounted
} the same way and placed in a vacuum? I wonder if their might be something
} galvanic going on with all the metals involved. Maybe the beam had nothing
} to do with it.
}
} Warren
}
} At 04:49 PM 05/29/05, you wrote:
}
}
} } ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 11:44:10 2005



From: Terry Mehr :      terry-at-AsylumResearch.com
Date: Mon, 6 Jun 2005 09:43:28 -0700
Subject: [Microscopy] AFM TechnicalSales Manager, UK Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Asylum Research is seeking a Technical Sales Executive to manage its
UK sales operations. The candidate should have experience both in
sales and technical applications for scanning probe/atomic force
microscopy (SPM/AFM) or scientific instrumentation/metrology tools.
Additional information on the position may be found at http://
www.AsylumResearch.com.

Terry Mehr
Director of Marketing
Asylum Research
www.AsylumResearch.com




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 11:54:28 2005



From: Ben McMorran :      mcmorran-at-physics.arizona.edu
Date: Mon, 6 Jun 2005 09:53:46 -0700
Subject: [Microscopy] Documentation for ISI WB6

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello,

We have scavenged the optics column from a defunct ISI WB6 electron microscope
(tungsten filament) in order to build an electron interferometer. Does anybody
out there have any technical specifications, engineering drawings, electrical
schematics, etc. for this equipment? We already have the user manual and an
adjustment manual.

Specifically, while we have been able to roughly focus and scan the beam using
our own power supplies, it would be really nice to know, for example, coil
currents that correspond to given focal lengths from a particular lens. Since
we're doing everything "by hand", we would love to know where all the beam
focal points are located. Dunno if that information is publicly available, but
I thought I'd ask. Thanks for your time!

--Ben McMorran

--
Ben McMorran
Research Assistant, Atom Optics Group
Department of Physics
University of Arizona
1118 E 4th St
Tucson, AZ
USA

ph. 520-621-2688



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 17:37:09 2005



From: Christine Weaver, McRI :      cweaver-at-mcri.org
Date: Mon, 6 Jun 2005 17:35:05 -0500
Subject: [Microscopy] Introduction to the Techniques of Forensic Soil Comparison Workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

"Introduction to the Techniques of Forensic Soil Comparison"
Workshop Conducted by Skip Palenik
Inter/Micro 2005
McCrone Research Institute
July 14-15, 2005

This 2-day workshop, conducted by Skip Palenik, renowned expert in the field
of forensic microscopy and microanalysis, introduces innovative techniques
for the examination and interpretation of soil samples and soil comparison
evidence. Using a combination of several techniques (including visual and
microscopical methods, filtration, ultra-sonication, sieving, and chemical
treatments) soil samples can be separated into useful fractions. You will
learn and practice Skip's modified fractionation techniques to investigate
the most distinguishing particles that might ordinarily be overlooked,
present in low quantities, or hidden by clay and humic materials. This
workshop is unique in that it emphasizes a microscopical approach on the
level at which soil is most fundamentally distinct, i.e., on a
particle-by-particle basis. Forensic scientists from around the world have
verified that the use of these techniques on soil evidence has helped them
to identify the original source of an unknown sample in addition to
facilitating the process of comparison of known and unknown samples from
almost any geographical location.

The workshop will be conducted in the laboratories and classrooms of McCrone
Research Institute in Chicago, Illinois. The cost of the workshop is $200
which includes the use of equipment, teaching materials, lunch, snacks, and
refreshments. Please register early http://mcri.org/IM_03homep.html, class
size is limited.

Regards,
Christine

Christine Weaver
Inter/Micro Coordinator
McCrone Research Institute
2820 S. Michigan Avenue
Chicago IL 60616

312-842-7100
cweaver-at-mcri.org
http://www.mcri.org





From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 18:26:56 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 06 Jun 2005 16:24:13 -0700
Subject: [Microscopy] RE: [MICROSCOPY] specimen heating up

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi all:

Problem uncovered. Reason why is not totally clear.

The situation is that if the colloidal silver is applied
and then the specimen is put in the SEM after about two to
three hours, it gets hot. Leave the specimen in air or
out for a day, put in SEM, no heat. OK. So, what seems
to be happening is that the binders in the colloidal silver
product outgasses rapidly in a vacuum and accelerates an
otherwise slow and un-noticed heating when done in air over
a long period of time.

So, the solution is to either vacuum dessicate the stub
before putting in the SEM or wait over night before putting
into the SEM.

Any other explanations?

gary g.





At 09:54 AM 6/6/2005, you wrote:
} Well, it seems to me that you can't have a sample heating up that much
} without the stage getting hot too. And you'd need a fair bit of
} electrical current dissipating in resistive heat to cause it, must be
} several hundred mA (depending how much material is getting hot). Seems
} like you have a hot stage. Perhaps a dodgy motor on the stage drive, or
} heat transmitted from a diff pump (can't see how though). Or friction,
} does it depend how much you move the stage about? I guess the stage is
} warm to the touch?
}
} When you say:
}
} 'After some amount of time in air, the specimens seem to not
} heat up. What is going on?'
}
} Do you mean the specimen has been in air for a while, you put it back in,
} and it doesn't get hot - or do you mean when the chamber is vented up to
} air, the specimen cools down. (I guess the latter, cooling by convection
} currents in the air then.)
}
} So I guess I would look for electrical current going into the stage (eg
} pull the plug from the stage drive controller board, see if the effect
} goes away), friction, or something connected to the stage that is warm..
} Interesting one though, I'd be intrigued to know what it turns out to be.
}
} Good luck
}
} Richard
}
}
} ________________________________________
} Richard Beanland
} Analytical Services
} Bookham Inc
} Caswell
} Towcester
} Northants
} NN12 8EQ
} United Kingdom
} Tel. +44 1327 356362
} Fax. +44 1327 356775
} http://www.bookham.com
} ________________________________________
}
}
} -----Original Message-----
} From: Gary Gaugler [mailto:gary-at-gaugler.com]
} Sent: 06 June 2005 17:24
} To: MSA listserver
} Subject: [Microscopy] [MICROSCOPY] specimen heating up
}
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 18:43:28 2005



From: Gordon Vrololjak :      gvrdolja-at-nature.berkeley.edu
Date: Mon, 6 Jun 2005 16:42:41 -0700 (PDT)
Subject: [Microscopy] Re: Nikon Fluophot

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

For manuals, contact Nikon directly (email/phone) and they will send you a
manual or a copy of one.
Gordon

\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\
Gordon Ante Vrdoljak Electron Microscope Lab
AOL/IM rakhasha http://nature.berkeley.edu/~gvrdolja 26 Giannini Hall
gvrdolja-at-nature.berkeley.edu UC Berkeley
phone (510) 642-2085 Berkeley CA 94720-3330
fax (510) 643-6207 cell (510) 290-6793

On Sat, 4 Jun 2005, km602223-at-comcast.net wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Does anyone have an instruction manual for a Nikon Fluophot microscope?
} Thank you,
} Kathleen
}


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 20:02:51 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 06 Jun 2005 17:58:57 -0700
Subject: [Microscopy] RE: [MICROSCOPY] specimen heating up

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Who said that laws cannot be broken? Speeders do it
all the time. Red light runners...etc. Let's give
someone a ticket!

All I can say is that this is what I see. Try the experiment
for yourself and see what you get. perhaps there is a
variable in this situation that I am missing.

gary g.


At 06:14 PM 6/6/2005, you wrote:
} Huh? You've uncovered a violation of the second law of thermodynamics
} where the
} transition liq } gas, or adsorbed specie } gas is exothermic in spite of the
} increase in free energy????
}
} John
}
} Gary Gaugler wrote:
}
} }
} ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
} -------------------------------------------------------------------------------
} }
} } Hi all:
} }
} } Problem uncovered. Reason why is not totally clear.
} }
} } The situation is that if the colloidal silver is applied
} } and then the specimen is put in the SEM after about two to
} } three hours, it gets hot. Leave the specimen in air or
} } out for a day, put in SEM, no heat. OK. So, what seems
} } to be happening is that the binders in the colloidal silver
} } product outgasses rapidly in a vacuum and accelerates an
} } otherwise slow and un-noticed heating when done in air over
} } a long period of time.
} }
} } So, the solution is to either vacuum dessicate the stub
} } before putting in the SEM or wait over night before putting
} } into the SEM.
} }
} } Any other explanations?
} }
} } gary g.
} }
} } At 09:54 AM 6/6/2005, you wrote:
} } } Well, it seems to me that you can't have a sample heating up that much
} } } without the stage getting hot too. And you'd need a fair bit of
} } } electrical current dissipating in resistive heat to cause it, must be
} } } several hundred mA (depending how much material is getting hot). Seems
} } } like you have a hot stage. Perhaps a dodgy motor on the stage drive, or
} } } heat transmitted from a diff pump (can't see how though). Or friction,
} } } does it depend how much you move the stage about? I guess the stage is
} } } warm to the touch?
} } }
} } } When you say:
} } }
} } } 'After some amount of time in air, the specimens seem to not
} } } heat up. What is going on?'
} } }
} } } Do you mean the specimen has been in air for a while, you put it back in,
} } } and it doesn't get hot - or do you mean when the chamber is vented up to
} } } air, the specimen cools down. (I guess the latter, cooling by convection
} } } currents in the air then.)
} } }
} } } So I guess I would look for electrical current going into the stage (eg
} } } pull the plug from the stage drive controller board, see if the effect
} } } goes away), friction, or something connected to the stage that is warm..
} } } Interesting one though, I'd be intrigued to know what it turns out to be.
} } }
} } } Good luck
} } }
} } } Richard
} } }
} } }
} } } ________________________________________
} } } Richard Beanland
} } } Analytical Services
} } } Bookham Inc
} } } Caswell
} } } Towcester
} } } Northants
} } } NN12 8EQ
} } } United Kingdom
} } } Tel. +44 1327 356362
} } } Fax. +44 1327 356775
} } } http://www.bookham.com
} } } ________________________________________
} } }
} } }
} } } -----Original Message-----
} } } From: Gary Gaugler [mailto:gary-at-gaugler.com]
} } } Sent: 06 June 2005 17:24
} } } To: MSA listserver
} } } Subject: [Microscopy] [MICROSCOPY] specimen heating up
} } }
} } }
} } }
} } }
} } } -----------------------------------------------------------------------
} -------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} } } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------
} --------
} } }
} } } I tried two other sets of specimens and they all heated up.
} } } The ambient temperature is 70F. The heated temperature is
} } } 95F.
} } }
} } } One specimen is a sputter coated (Pt) #1 cover slip stuck on
} } } a pin stub via Nisshin EMS carbon tab. The slip is offset
} } } slightly so I can ground it using Pella colloidal silver.
} } }
} } } When I put this in the chamber, no beam, for about 10 minutes,
} } } I pull it out and it is hot.
} } }
} } } Then I did three Al film on silicon pieces on stubs which were
} attached using
} } } Buehler wax but grounded using the Pella colloidal silver.
} } } This large holder also got up to 95F in chamber vacuum without
} } } any beam.
} } }
} } } After some amount of time in air, the specimens seem to not
} } } heat up. What is going on?
} } }
} } } Strange. But it messes up image collection due to large
} } } amount of drift.
} } }
} } } gary g.
} } }
} } }
} } }
} } } At 08:30 AM 5/31/2005, you wrote:
} } } } Just a wild guess, but would the sample have heated up the same if
} mounted
} } } } the same way and placed in a vacuum? I wonder if their might be
} something
} } } } galvanic going on with all the metals involved. Maybe the beam had
} nothing
} } } } to do with it.
} } } }
} } } } Warren
} } } }
} } } } At 04:49 PM 05/29/05, you wrote:
} } } }
} } } }
} } } } } --------------------------------------------------------------------
} ----
} } } ------
} } } } } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} } } } } To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} } } } } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } } } } --------------------------------------------------------------------
} ----
} } } -------
} } } } }
} } } } } Hi Listers:
} } } } }
} } } } } I've encountered a once only situation that I cannot
} } } } } explain. Perhaps someone has some ideas about what
} } } } } is going on.
} } } } }
} } } } } I put a specimen in the SEM for examination and when
} } } } } I pulled it out via the load lock, the specimen and
} } } } } holder were about 10 degrees hotter than ambient.
} } } } } My experience has been that whatever they go in at,
} } } } } they come out the same. This was simple SE imaging
} } } } } at 10KV with 30u High Current in Zeiss Supra 55VP
} } } } } in high vacuum mode. Specimen current was around 145pA.
} } } } }
} } } } } The specimen was a copper tube that had been plated
} } } } } with silver on the outside and inside. The piece was
} } } } } one half of the tube and about .5" long, attached to
} } } } } a pin stub using colloidal silver.
} } } } }
} } } } } EDS showed presence of K, O, Ag, Cu, and C. There were
} } } } } distinct areas of organic material that I take to be
} } } } } cyanide from the plating process. But how could SEM
} } } } } analysis of a specimen cause it to heat?
} } } } }
} } } } } Any ideas?
} } } } }
} } } } } gary g.
} } }
} } }
} } }
} } } =======================================================================
} } } This e-mail is intended for the person it is addressed to only. The
} } } information contained in it may be confidential and/or protected by
} } } law. If you are not the intended recipient of this message, you must
} } } not make any use of this information, or copy or show it to any
} } } person. Please contact us immediately to tell us that you have
} } } received this e-mail, and return the original to us. Any use,
} } } forwarding, printing or copying of this message is strictly prohibited.
} } } No part of this message can be considered a request for goods or
} } } services.
} } } =======================================================================



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 21:21:21 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 06 Jun 2005 18:31:56 -0700
Subject: [Microscopy] Re: RE: [MICROSCOPY] specimen heating up

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I tried three pieces of Si coated with colloidal conductive
silver and began dessicating them in a Duniway chamber. This
got down to about 50mT using a Varian dual vane mech pump.

After thirty minutes, the Zeiss holder is not hot. So,
next is moving it into the E-6 Torr SEM chamber. More to
follow.

gary g.



At 05:58 PM 6/6/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 22:29:04 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Mon, 06 Jun 2005 20:24:49 -0700
Subject: [Microscopy] Re: [MICROSCOPY] specimen heating up

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Well, I think this problem is solved/isolated. The laws of
thermo are preserved. The SEM is a Zeiss Supra 55VP.
It has an Edwards XDS10 scroll pump and Varian TMP.
It is a dry system. Specimens are introduced via
a Fjeld load lock. All together, they are great.
The current problem seems to be the stage rotate motor.

It turns out that depending on some variables, the
Zeiss Supra 55VP stage rotate motor gets very hot.
This translates to a hot specimen holder and hot
specimens.

I see no way to avoid this other than to turn off R
until a new specimen needs it. This is not optimum.
But it seems to be the only solution. There were many
start up problems with this unit and the stage was
one of them. R was a big issue. I guess that it
has not gone away.

Thus, colloidal silver is not the culprit/cause.

gary g.


At 06:31 PM 6/6/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 6 23:09:30 2005



From: Nestor J. Zaluzec :      zaluzec-at-microscopy.com
Date: Mon, 6 Jun 2005 23:08:49 -0500
Subject: [Microscopy] Administrivia - correction for latest Microscopy Today

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} To: DrJohnRuss-at-aol.com
} Subject: [Microscopy] RE: correction for latest Microscopy Today

John

I have read with interest your recent Listserver posting of a few days ago. There is useful information in your text, specifically your methodology within Photoshop tools to construct stereo pairs. However, I note with concern that the first paragraph of your posting contains content which is in violation of the Microscopy Listserver rules.

This concern is not something to be taken lightly and I will quote from the Microscopy Listserver rules, which as subscribers EVERYONE receives a copy of (http://www.microscopy.com/Rules.html), and is a condition of their continued subscription. Specifically item #4 of those rules states:

----------------------------------------------------------------------
4.) This forum is not to be used as a platform to accuse or defame any individual or organization in any negative manner. You may disagree with any comment posted and post a reply, but this server may not be used to spread misleading, derogatory or disparaging comments under any conditions.
----------------------------------------------------------------------

In my opinion, you have crossed the line with your initial comments. It is fine to ask questions and/or offer alternatives or comments about any posting on the Listserver as that is certainly one of the purposes of this forum. However, in your posting you are making comments to the Listserver about an article which has no relation to the Listserver, the text for this technique never appeared in the Listserver or its archives, nor has anyone asked a question concerning this technique on the Listserver. Using this forum to disparage anyone is against our rules, and both Microscopy-Today and Jerry Sedgewick are no different than anyone else. I have not stood by idly when comparable situations happened in the past, nor will I do so in this situation.

I reiterate, posting information on your technique to create stereo pairs using Photoshop is fine as well as offering it as an alternative to methods of others. It is also fine that you disagree with Sedgewick (I disagree with alot of people and have disagreed with you on various issues over the years). Since you feel so strongly about this particular article in MT, I would encourage you to write a contributed article and submit it to MT about the methodoloy you just outlined, however, using the Listserver is an inappropriate forum to voice complaints or views about MT and/or Sedgewick.

If you (or anyone for that matter) feel that you are being ignored or have problems with MSA or one of its subsidiaries, please feel free to contact me off-line to discuss the situation and please send me a copy of any documentation you have submitted to the parties in question. As a former Council member I will be willing to bring your concerns to the attention of the appropriate individuals within MSA, or to the Council as appropriate.

I insist that EVERYONE using the Microscopy Listserver follows proper channels when they have problems with other organizations or individuals (MSA related or not) and no one is permitted to use the Listserver as a forum to disparage any individual or organization.


Nestor
Your Friendly Neighborhood SysOp
&
The Microscopy Listserver SysCop


-------------------------------------------------------------------------------
} From: DrJohnRuss-at-aol.com
} Date: Sat, 4 Jun 2005 09:19:58 EDT
} Subject: [Microscopy] correction for latest Microscopy Today
} To: microscopy-at-msa.microscopy.com


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
--------cut-----------




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 10:29:57 2005



From: Mike Zemyan :      mzemyan-at-schaferlabs.com
Date: Tue, 7 Jun 2005 08:29:08 -0700
Subject: [Microscopy] Mass Spectrometry Engineer / Technician Job Opening at Schafer Corporation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Mass Spectrometry Engineer / Technician Job Opening
at Schafer Corporation

Schafer Corporation is seeking an engineer / technician for its Mass
Spectrometry Group at its Vallecitos Laboratory near Sunol, California.
This position supports a multi-million dollar operation of thermal
ionization and secondary ion mass spectrometers and support instruments, and
interacts with a team of mass spectrometry scientists, engineers and
technicians. The candidate will also be expected to directly participate in
the analysis of samples and in the maintenance of auxiliary equipment.
Schafer Vallecitos Laboratory performs materials characterization and
related analytical services on commercial and government contracts. The
activities of the group include chemical analysis of materials on a
production basis, maintenance of several custom-built mass spectrometers,
development of new or improved mass spectrometric techniques, and quality
assurance of analytical data.

This position requires an individual who has proven technical abilities.
The ideal candidate must have a strong background in science or engineering,
or a related field of physical or chemical science, and experience operating
mass spectrometers or other analytical equipment in a laboratory
environment. The candidate must be able to operate independently with
little supervision. Other desired qualifications include:

- Expertise in data evaluation and quality control.
- Ability to write clear scientific and technical reports.
- Bachelor's degree in physical science or engineering.
- Minimum 4 years of laboratory experience.

Schafer offers a comprehensive array of benefits including dental, health
and life insurance, 4 weeks of vacation annually, and a 401k program. US
Citizenship required, as well as qualification for a DoD security clearance
- An Affirmative Action, EEO employer promoting diversity in the workplace.

Respond to job code VAL200502 - at jobs2-at-schaferlabs.com






From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 11:17:29 2005



From: John J. Bozzola :      bozzola-at-siu.edu
Date: Tue, 7 Jun 2005 14:01:02 -0500
Subject: [Microscopy] high pass filter versus flat field correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Skage,
There are a few other possibilities that come to mind. First, you sample may
be grounded to the stub, but have you traced the ground all the way? Check
the sample holder, stage and the wire that goes to the feedthrough. My stage
grounding goes through a BNC plug, in case you want to test the specimen
current, and sometimes the plug needs cleaning because the oxide film builds
up and stops the grounding current. Remember the current is picoamps, so it
doesn't take much to stop it. I have had the grounding wire in the chamber
broken because it got caught in the o-ring when the chamber was pumped down.
The other thing to check is the scintillator. The aluminum coating on the
scintillator can become worn off by micro-discharges, until the bare plastic
button is showing, and this can build up a charge. A new scintillator button
is a good idea, or a new evaporated layer of Al and check that the button is
well grounded.
Of course dirt in the column or stage can cause these problems, as well.
----- Original Message -----
} From: "Skage Hem" {SHem-at-laurentian.ca}
To: {Microscopy-at-microscopy.com}
Sent: Monday, June 06, 2005 8:29 AM

I am working with digital TEM images that need flat field correction
and background subtraction to reduce fixed, uneven illumination and
noise (lint & electronic noise), respectively. I have a program that
came with the digital camera but it is cumbersome to use (at least,
so far).

In PhotoShop (version 6 and CS1) I can flatten the field using a high
pass filter but have been unable to use PS to subtract noise
(original image minus empty field). In PS, I tried using the
Image/Calculations feature as well as Image/Apply Image features. No
combination gave the desired effects.

Am I asking too much of PhotoShop, or simply not understanding how to
achieve this in PS? Maybe both, eh?

Any suggestions would be appreciated.

Thank you.
--
##############################################################
John J. Bozzola, Ph.D., Director
I.M.A.G.E. (Integrated Microscopy & Graphics Expertise)
750 Communications Drive - MC 4402
Southern Illinois University
Carbondale, IL 62901 U.S.A.
Phone: 618-453-3730
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/~image/
##############################################################


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 14:10:51 2005



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Tue, 07 Jun 2005 14:10:08 -0500
Subject: [Microscopy] Sectioning stainless steel wire

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

This is for the material Scientists among you,

We need to make 20-30µm sections of 200µm thick stainless steel wire coated
with a polymer. The object is to be able to do some chemical analysis on
the sections so as to visualize chemical distribution in the polymeric
coating. The sections will be imaged using Ramon technology.

We need suggestions as to embedding resin, type of microtome and type of
blades that may work for sections of this thickness and samples of this
type.

Thanks,
Debby

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy





From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 14:30:04 2005



From: DrJohnRuss-at-aol.com
Date: Tue, 7 Jun 2005 15:29:15 EDT
Subject: [Microscopy] Re: high pass filter versus flat field correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

There are three classic approaches to levelling background. Some are easier
to do in Photoshop than others.
1) If you can measure a background (ie. collect an image with no sample, or a
uniform sample present) then you can use layers modes to get the difference
between the original and the background. Unfortunately, depending on how your
camera works you may need to either subtract the background or divide by it (if
the camera is log or linear, respectively). Photoshop doesn't offer a ratio
layers mode (these are the same options as in the Calculations dialog), so you
would have to use a plugin (e.g., the Reindeer Fovea Pro suite which can
either subtract or divide.
2) If you can remove the features from the image (e.g. if they are skinny in
one dimension) to leave a background representation, then you can use that. In
Photoshop you can do a slightly crude but probably adequate job of a grey
scale morpholological closing or opening using the Filter-} Custom-} Maximum or
Minimum routine. For example if the features are dark on a bright background, you
would first use Maximum to remove the features and then Minimum with the same
radius to restore the scale of the background irregularities. This background
is then removed as in step 1.
3) In many cases, the gradually varying background from nonuniform
illumination can be modelled mathematically. In Fovea Pro, a cubic polynomial is fitted
to either the bright or dark parts of the image to construct a function for
removal. Sometimes you can accomplish the same thing with a large low-pass
filter (e.g., a Gaussian smooth). This blurs out the features to the point where
the image shows just the background, which you can then use as in step 1.

John Russ
=======

In a message dated 6/7/05 3:03:50 PM, bozzola-at-siu.edu writes:

} I am working with digital TEM images that need flat field correction
} and background subtraction to reduce fixed, uneven illumination and
} noise (lint & electronic noise), respectively. I have a program that
} came with the digital camera but it is cumbersome to use (at least,
} so far).
}
} In PhotoShop (version 6 and CS1) I can flatten the field using a high
} pass filter but have been unable to use PS to subtract noise
} (original image minus empty field). In PS, I tried using the
} Image/Calculations feature as well as Image/Apply Image features. No
} combination gave the desired effects.
}
} Am I asking too much of PhotoShop, or simply not understanding how to
} achieve this in PS? Maybe both, eh?
}
} Any suggestions would be appreciated.


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 14:44:34 2005



From: Roger Baker :      rcbaker-at-eden.infohwy.com
Date: Tue, 7 Jun 2005 14:43:42 -0500
Subject: [Microscopy] Re: Sectioning stainless steel wire

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I suspect that slicing through stainless steel wire without
disturbing its polymer coating on a microscopic level is going to be
a huge challenge.

The hard wire would probably mess up knives and if you try to grind a
flat on the end of the wire, the abrasive will get embedded in the
interface.

I would suggest simply slicing/stripping off the coating and looking
at that in cross section since its surface chemistry is not going to
change with the removal of the wire in contact.

-- Roger


On Jun 7, 2005, at 2:10 PM, Debby Sherman wrote:

}
}
} ----------------------------------------------------------------------
} --------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/
} MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} ----------------------------------------------------------------------
} ---------
}
} This is for the material Scientists among you,
}
} We need to make 20-30µm sections of 200µm thick stainless steel
} wire coated
} with a polymer. The object is to be able to do some chemical
} analysis on
} the sections so as to visualize chemical distribution in the polymeric
} coating. The sections will be imaged using Ramon technology.
}
} We need suggestions as to embedding resin, type of microtome and
} type of
} blades that may work for sections of this thickness and samples of
} this
} type.
}
} Thanks,
} Debby
}
} Debby Sherman, Manager Phone: 765-494-6666
} Life Science Microscopy Facility FAX: 765-494-5896
} Purdue University E-mail: dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907
} http://www3.agriculture.purdue.edu/microscopy
}
}
}
}
}
}




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 15:59:40 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Tue, 7 Jun 2005 13:58:53 -0700
Subject: [Microscopy] Sectioning stainless steel wire

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Debbie;
FIB.

John Mardinly
Intel


-----Original Message-----
} From: Debby Sherman [mailto:dsherman-at-purdue.edu]
Sent: Tuesday, June 07, 2005 12:10 PM
To: message to: MSA list

This is for the material Scientists among you,

We need to make 20-30µm sections of 200µm thick stainless steel wire coated
with a polymer. The object is to be able to do some chemical analysis on
the sections so as to visualize chemical distribution in the polymeric
coating. The sections will be imaged using Ramon technology.

We need suggestions as to embedding resin, type of microtome and type of
blades that may work for sections of this thickness and samples of this
type.

Thanks,
Debby

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy







From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 16:09:47 2005



From: Thomas, Larry (PNNL) :      Larry.Thomas-at-pnl.gov
Date: Tue, 07 Jun 2005 14:09:04 -0700
Subject: [Microscopy] RE: high pass filter versus flat field correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Here's a favorite Photoshop recipe for background leveling digital TEM images (until someone suggests a better one). It uses Gaussian Blur as mentioned by J. Russ.

1. Open image or paste into new PS window.
2. Duplicate layer (twice)
3. Apply Gaussian Blur (50 pixels or your choice) to top layer
4. Invert image
5. Overlay image (blending mode option from Layers window)
6. Merge down

Repeat leveling operation 2 to 4 times until artifacts appear or contrast washes out.
Record as Action with different blur settings of choice.

For a more inclusive view of background leveling, see Russ tutorial at
reindeergraphics.com/tutorial/chap2/defect06.html

Larry Thomas
} ----------
} From: John J. Bozzola
} Sent: Tuesday, June 7, 2005 12:01 PM
} To: Microscopy-at-msa.microscopy.com
} Subject: [Microscopy] high pass filter versus flat field correction
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} I am working with digital TEM images that need flat field correction
} and background subtraction to reduce fixed, uneven illumination and
} noise (lint & electronic noise), respectively. I have a program that
} came with the digital camera but it is cumbersome to use (at least,
} so far).
}
} In PhotoShop (version 6 and CS1) I can flatten the field using a high
} pass filter but have been unable to use PS to subtract noise
} (original image minus empty field). In PS, I tried using the
} Image/Calculations feature as well as Image/Apply Image features. No
} combination gave the desired effects.
}
} Am I asking too much of PhotoShop, or simply not understanding how to
} achieve this in PS? Maybe both, eh?
}
} Any suggestions would be appreciated.
}
} Thank you.
} --
} ##############################################################
} John J. Bozzola, Ph.D., Director
} I.M.A.G.E. (Integrated Microscopy & Graphics Expertise)
} 750 Communications Drive - MC 4402
} Southern Illinois University
} Carbondale, IL 62901 U.S.A.
} Phone: 618-453-3730
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/~image/
} ##############################################################
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 16:12:32 2005



From: Bobby Hooghan :      hooghan-at-grandecom.net
Date: Tue, 7 Jun 2005 16:11:43 -0500
Subject: [Microscopy] Sectioning stainless steel wire

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Debbie,
Second John's suggestion whole heartedly, you can lift out a small sample
with no trouble at all, let me know if you need help,
Bobby Hooghan,


-----Original Message-----
} From: Mardinly, John [mailto:john.mardinly-at-intel.com]
Sent: Tuesday, June 07, 2005 3:59 PM
To: Debby Sherman; message to: MSA listDebbie,

Debbie;
FIB.

John Mardinly
Intel


-----Original Message-----
} From: Debby Sherman [mailto:dsherman-at-purdue.edu]
Sent: Tuesday, June 07, 2005 12:10 PM
To: message to: MSA list

This is for the material Scientists among you,

We need to make 20-30µm sections of 200µm thick stainless steel wire coated
with a polymer. The object is to be able to do some chemical analysis on
the sections so as to visualize chemical distribution in the polymeric
coating. The sections will be imaged using Ramon technology.

We need suggestions as to embedding resin, type of microtome and type of
blades that may work for sections of this thickness and samples of this
type.

Thanks,
Debby

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy










From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 18:35:54 2005



From: Connie McManus :      conniemoss-at-relia.net
Date: Tue, 7 Jun 2005 17:35:07 -0600 (MDT)
Subject: [Microscopy] Re: RE: Sectioning stainless steel wire

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

It's no lie that FIB would be the best solution to this problem, but what
if you don't have a FIB?? It is actually easier to cut metals very thin
i.e. 20-80 nm. At this thickness, they can be examined in the EFTEM,
which will analyze and map the elemental composition at high resolution.


Bill McManus
Mt Ogden Scientific Services -Providing full service electron microscopy
for you
950 W Kershaw, Suite E
Ogden UT 84401
toll-free: 877/311-6677
direct: 801/745-2583
cell: 435/757/2975
fax: 435/514-1781
conniemoss-at-relia.net
www.mtogdensci.com


-------------------------------
Mardinly, John said:
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Debbie;
} FIB.
}
} John Mardinly
} Intel
}
}
} -----Original Message-----
} } From: Debby Sherman [mailto:dsherman-at-purdue.edu]
} Sent: Tuesday, June 07, 2005 12:10 PM
} To: message to: MSA list
} Subject: [Microscopy] Sectioning stainless steel wire
}
}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} This is for the material Scientists among you,
}
} We need to make 20-30µm sections of 200µm thick stainless steel wire
} coated
} with a polymer. The object is to be able to do some chemical analysis on
} the sections so as to visualize chemical distribution in the polymeric
} coating. The sections will be imaged using Ramon technology.
}
} We need suggestions as to embedding resin, type of microtome and type of
} blades that may work for sections of this thickness and samples of this
} type.
}
} Thanks,
} Debby
}
} Debby Sherman, Manager Phone: 765-494-6666
} Life Science Microscopy Facility FAX: 765-494-5896
} Purdue University E-mail: dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907
} http://www3.agriculture.purdue.edu/microscopy
}
}
}
} c
}
}
}




======


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 21:05:12 2005



From: sdunning-at-uga.edu (by way of Ask-A-Microscopist)
Date: Tue, 7 Jun 2005 21:04:31 -0500
Subject: [Microscopy] [Filtered] AskAMicroscopist: how to calculate AFM resolution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (sdunning-at-uga.edu) from http://www.microscopy.com/Ask-A-Microscopist/Ask-A-Microscopist.html on Tuesday, June 7, 2005 at 10:57:01
---------------------------------------------------------------------------

Email: sdunning-at-uga.edu
Name: Sarah Dunning

Organization: University of Georgia

Education: Graduate College

Location: Athens, Georgia

Question: I read in the literature that AFM's achieve "atomic resolution." How does one go about calculating this mathematically? Is it a matter of the tip size, geometry, tip-sample sensitivity, etc., all folded into one equation?



---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 21:06:37 2005



From: przybylowicz-at-tlabs.ac.za (by way of MicroscopyListserver)
Date: Tue, 7 Jun 2005 21:05:54 -0500
Subject: [Microscopy] [Filtered] MicroscopyListserverviaWWW: Search for second hand

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (przybylowicz-at-tlabs.ac.za) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Monday, June 6, 2005 at 12:44:43
---------------------------------------------------------------------------

Email: przybylowicz-at-tlabs.ac.za
Name: Wojciech Przybylowicz

Organization: Materials Research Group, iThemba LABS

Title-Subject: [Microscopy] [Filtered] Search for the second hand automatic refilling system

Question: Our laboratory is looking for the automatic refilling device (7019-11006) that operates with Reichert-Jung FC4 Ultracut Cryosystem. It is composed of the control module (6527-00001), dewar with solenoid valve (7019-31030) and heating element inside, and cooling chamber (7019-1150). If you know of any laboratory willing to offer it to us for free or for a reasonable price, please let me know.

Wojciech Przybylowicz

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 7 21:07:20 2005



From: kristinat-at-hotmail.com (by way of MicroscopyListserver)
Date: Tue, 7 Jun 2005 21:06:36 -0500
Subject: [Microscopy] [Filtered] MicroscopyListserverviaWWW: UV filter for a 96-well

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (kristinat-at-hotmail.com) from http://www.microscopy.com/MLFormMail.html on Tuesday, June 7, 2005 at 11:09:35
---------------------------------------------------------------------------

Email: kristinat-at-hotmail.com
Name: Kristina Turner

Organization: New Mexico State University

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I am looking for a 620 or 625 nm UV filter for a 96-well plate reader. I will get the details of model number, etc. of the plate reader it would need to fit. I am doing Minimal Inhibitory Concentration susceptibility testing on Staphylococcus aureus with gossypol for my Master's thesis research project. The Professor in the Biology department that has a 96-well plate reader that I could use to read my results does not have a 625 nm UV filter that I need.
His plate reader is 10 years old, so it is pretty old.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 00:43:47 2005



From: Gordon Couger :      gcc-at-couger.com
Date: Wed, 08 Jun 2005 00:43:03 -0500
Subject: [Microscopy] Re: Re: high pass filter versus flat field correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


ImageJ http://rsb.info.nih.gov/ij/ is bit harder to use but will
do perfectly remove a background image from form an image by
taking a blank back ground image an using the
Process} math} difference to remove the blank frame from the
image. I leaves an image that need to be put on a flat back
ground but I haven't got a round toit yet.

ImageJ is a free tool developed by NIH and run on almost all
platforms. It has a large development community writing plug ins
and background is mentioned in several pulgins.

Gordon Couger
Stillwater, OK
www.couger.com/gcouger


} In a message dated 6/7/05 3:03:50 PM, bozzola-at-siu.edu writes:
}
}
} } I am working with digital TEM images that need flat field
correction
} } and background subtraction to reduce fixed, uneven
illumination and
} } noise (lint & electronic noise), respectively. I have a
program that
} } came with the digital camera but it is cumbersome to use (at
least,
} } so far).
} }
} } In PhotoShop (version 6 and CS1) I can flatten the field
using a high
} } pass filter but have been unable to use PS to subtract noise
} } (original image minus empty field). In PS, I tried using the
} } Image/Calculations feature as well as Image/Apply Image
features. No
} } combination gave the desired effects.
} }
} } Am I asking too much of PhotoShop, or simply not
understanding how to
} } achieve this in PS? Maybe both, eh?
} }
} } Any suggestions would be appreciated.
}
}
}




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 04:09:50 2005



From: Richard Beanland :      richard.beanland-at-bookham.com
Date: Wed, 8 Jun 2005 10:08:24 +0100
Subject: [Microscopy] Sectioning stainless steel wire

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I would have concerns about surface contamination from FIB which could affect the results, not to mention changes in the polymer itself from damage to the ion beam.
This isn't too different from the optoelectronic packages we section on occasion, which can have everything from stainless steel through glass to adhesives. It's not clear to me why you need a thin section for Raman - can't you just pot the wire in an appropriate resin (we use Struers but I'm sure other suppliers are just as good) and grind/polish as for a standard metallographic section? When there are changes in material hardness it can be difficult to keep the surface flat, but a little experimentation for your application with different pads should get you there. This should give you a nice clean and flat surface for analysis, and a few polishing pads and a bit of resin is slightly cheaper than a FIB.

Richard

________________________________________
Richard Beanland
Analytical Services
Bookham Inc
Caswell
Towcester
Northants
NN12 8EQ
United Kingdom
Tel. +44 1327 356362
Fax. +44 1327 356775
http://www.bookham.com
________________________________________


-----Original Message-----
} From: Debby Sherman [mailto:dsherman-at-purdue.edu]
Sent: 07 June 2005 20:10
To: message to: MSA list

This is for the material Scientists among you,

We need to make 20-30µm sections of 200µm thick stainless steel wire coated
with a polymer. The object is to be able to do some chemical analysis on
the sections so as to visualize chemical distribution in the polymeric
coating. The sections will be imaged using Ramon technology.

We need suggestions as to embedding resin, type of microtome and type of
blades that may work for sections of this thickness and samples of this
type.

Thanks,
Debby

Debby Sherman, Manager Phone: 765-494-6666
Life Science Microscopy Facility FAX: 765-494-5896
Purdue University E-mail: dsherman-at-purdue.edu
S-052 Whistler Building
170 S. University Street
West Lafayette, IN 47907
http://www3.agriculture.purdue.edu/microscopy





=======================================================================
This e-mail is intended for the person it is addressed to only. The
information contained in it may be confidential and/or protected by
law. If you are not the intended recipient of this message, you must
not make any use of this information, or copy or show it to any
person. Please contact us immediately to tell us that you have
received this e-mail, and return the original to us. Any use,
forwarding, printing or copying of this message is strictly prohibited.
No part of this message can be considered a request for goods or
services.
=======================================================================



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 06:26:11 2005



From: michael shaffer :      michael-at-shaffer.net
Date: Wed, 8 Jun 2005 08:55:29 -0230
Subject: [Microscopy] ESEM: accelerating oxidation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Is anyone aware of a method for accelerating oxidation while observing with
ESEM?

tia :o)
michael shaffer
(709) 737-8346 (voice)
(709) 737-2589 (FAX)
{www.micro-investigations.com}
{www.esd.mun.ca/epma/}
Alexander Murray Bldg, Rm ER4063
Memorial University of Newfoundland
St. John's, NL
Canada, A1B 3X5



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 06:31:53 2005



From: microbill-at-mohawk.net
Date: Wed, 8 Jun 2005 07:31:10 -0400
Subject: [Microscopy] Job Opening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



HIROX-USA, INC.
}
} Hirox USA, distributor of Industrial Video-Microscope systems, is seekng a
} Sales Engineer who has prior experience with the sale of capital equipment.
} The job requirements include: Demonstration of the instruments, travel and
} the ability to lift equipment up to 100LB.
} The position is for the New Jersey area. This industry is growing
} rapidly and our customers include many Fortune 500 companies. see
} {http://www.hirox-usa.com} www.hirox-usa.com Email to:
} {mailto:mail-at-hirox-usa.com} mail-at-hirox-usa.com or Fax resume to 201.342.7322


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 11:53:03 2005



From: Kestutis Smalinskas :      smalinskas-at-yahoo.com
Date: Wed, 8 Jun 2005 09:52:18 -0700 (PDT)
Subject: [Microscopy] Any SEM/EDS labs in Shanghai, China?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I just got a request from someone in China if there
are any labs available in China that can do SEM/EDS
work. Can anyone provide contact information?

Stu Smalinskas, P.E.
Metallurgist
SKF North American Technical Center
Plymouth, Michigan
(734) 414-6862
stu.smalinskas-at-skf.com



__________________________________
Discover Yahoo!
Have fun online with music videos, cool games, IM and more. Check it out!
http://discover.yahoo.com/online.html


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 13:34:07 2005



From: Kestutis Smalinskas :      smalinskas-at-yahoo.com
Date: Wed, 8 Jun 2005 11:32:52 -0700 (PDT)
Subject: [Microscopy] Re: Any SEM/EDS labs in Shanghai, China?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

To add, this request was from a manufacturer of
automotive mirrors. They need EDS work on the layers
of the mirrored surface. I’ve been doing this work
for their customer, but they want to have it done
locally for themselves.

Stu

--- Changhui LEI {clei-at-uiuc.edu} wrote:

} Contact Shanghai Jiaotong U, Fudan U.
}
} They should have good TEM, SEM, lab.
}
} Or you can contact SMICS. They have SEM/EDS. But I
} do not know if they like to check metals for you as
} SMICS is a semiconductor manufacture.
}
} Ch
}
} ---- Original message ----
} } Date: Wed, 8 Jun 2005 09:52:18 -0700 (PDT)
} } From: Kestutis Smalinskas {smalinskas-at-yahoo.com}
} } Subject: [Microscopy] Any SEM/EDS labs in Shanghai,
} China?
} }
} }
} } I just got a request from someone in China if there
} } are any labs available in China that can do SEM/EDS
} } work. Can anyone provide contact information?
} }
} } Stu Smalinskas, P.E.
} } Metallurgist
} } SKF North American Technical Center
} } Plymouth, Michigan
} } (734) 414-6862
} } stu.smalinskas-at-skf.com




__________________________________
Discover Yahoo!
Find restaurants, movies, travel and more fun for the weekend. Check it out!
http://discover.yahoo.com/weekend.html



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 13:37:50 2005



From: Scott Griffith :      skod-at-ises-llc.com
Date: Wed, 08 Jun 2005 12:37:07 -0600
Subject: [Microscopy] Wanted to buy: used/serviceable stepper stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I need a working/repairable scanning stage, ideally already set up for
Leitz Orthoplan or Ergolux-compatible mounting dimensions. I've
essentially abandoned trying to get my Ergolux's servo-driven stage
working again due to the unavailability of parts and diagnostic info. The
eventual goal is to drive the unit with an Oasis-4i to create an automated
image capture system. If you have applicable stage hardware available for
a reasonable price, please contact me. Thanks in advance.

--
Scott Griffith
ISES-LLC
9745 Steeplechase Drive
Franktown, CO 80116
303-660-2541 voice
303-660-2542 fax
skod-at-ises-llc.com


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 15:06:32 2005



From: Gib Ahlstrand :      ahlst007-at-umn.edu
Date: Wed, 08 Jun 2005 15:05:18 -0500
Subject: [Microscopy] Diffusion pump oils

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Listers,

I'm trying to characterize two kinds of diffusion pump oils that I found in
my "back room":

1. Balzers-Oil 71 für Diffusionspumpen. Ch 76-946


2. Trafo Olie, Dialac C #132250466801)


They both look like they have been around for some years. Does anyone know
what kind of oils they are, eg. Silicone based or not?

My TEM and SEM have dedicated diff pump oils supplied by the manufacturer,
but I'm wondering if either of the ones mentioned above would be OK to use
in the diff pump of my Denton vacuum evaporator, which I use for routine
carbon or gold-palladium coatings.

Thanks in advance!

Gib
--
Gib Ahlstrand, Scientist
Electron Optical Facility, University of Minnesota, CBS Imaging Center,
35 Snyder Hall, St. Paul, MN. USA. 55108 (612)624-3454
(612)624-2785 FAX, ahlst007-at-tc.umn.edu
http://www.cbs.umn.edu/ic/





From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 16:16:27 2005



From: Elaine Humphrey :      ech-at-interchange.ubc.ca
Date: Wed, 08 Jun 2005 14:58:57 -0700
Subject: [Microscopy] Re: hydrogel mesh

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

The second position (Trafo Olie, Dialac C) is a transformer oil- don't use
it in vac. pumps. It is cheap - under $4 per gallon. If it is old and
discolored- dispose of it. The first one (Balzers-Oil 71) must be OK for
diff. pump use but:

a) is it in good condition (clean? clear? not oxydized/contaminated)?

b) you must completely clean traces of previously used oil from diff pump
(these oils don't perform well if at all when mixed), unless you know for
sure that diff pump was previously filled with the same type oil;

c) unless mistery oil is Santovac or Pentavac (very viscous at room
temperature, like honey, clear, either colorless or has very light yellow
color), it is cheap and IMO isn't worth the effort of trying it with
uncertain results. But I would definitely try old Santovac, especially for a
large diffusion pump, since it costs $1 and more per cc. Otherwise don't
bother.

Vitaly Feihgold
SIA
2773 Heath Lane
Duluth, GA 30096
Ph. 770-232-7785
Fax 770-232-1791
www.sia-cam.com

----- Original Message -----
} From: "Gib Ahlstrand" {ahlst007-at-umn.edu}
To: "Microscopy Listserver" {Microscopy-at-MSA.Microscopy.com}
Sent: Wednesday, June 08, 2005 4:05 PM

Hi Pat
I agree with Debby. We have one too but I think your nearest cryo sem
is University of New Brunswick, Frederickton.
Elaine

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
Dr. Elaine Humphrey
Director, BioImaging Facility
President, Microscopy Society of Canada (2003-2005)
University of British Columbia
6270 University Blvd, mail-stop Botany
Vancouver, BC
CANADA, V6T 1Z4
Phone: 604-822-3354
FAX: 604-822-6089
e-mail: ech-at-interchange.ubc.ca
website: www.emlab.ubc.ca


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 17:05:13 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Wed, 08 Jun 2005 15:04:48 -0700
Subject: [Microscopy] Re: Re: Diffusion pump oils

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Vitaly right
I would not put unknown quality diff oil into my pump: if it went wrong,
you will spend days to clean pump up. Personally, I am using Santovac -
it's serving my DV502A for 11 years without any problems: vacuum 5x10-7
torr after a few hours, no LN2. Sergey

At 01:58 PM 6/8/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu






From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 20:14:48 2005



From: acassell-at-mail.arc.nasa.gov (by way of MicroscopyListserver)
Date: Wed, 8 Jun 2005 20:14:05 -0500
Subject: [Microscopy] viaWWW: video capture card for Gatan Camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (acassell-at-mail.arc.nasa.gov) from
http://www.microscopy.com/MLFormMail.html on Wednesday, June 8, 2005
at 19:04:33
---------------------------------------------------------------------------

Email: acassell-at-mail.arc.nasa.gov
Name: Alan Cassell

Organization: NASA Ames Research Center

Title-Subject: [Microscopy] [Filtered] MListserver: Gatan video capture card?

Question: We have a bottom mount Gatan fiber optic coupled camera on
our JEOL2000FX TEM and we would like to use a video capture card to
grab frames from the monitor for the camera. Has anyone tried
regular commercial video cards? And if so, which ones work the best?

Thanks,

Alan

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 8 20:45:51 2005



From: Anthony Garratt-Reed :      tonygr-at-MIT.EDU
Date: Wed, 08 Jun 2005 21:41:36 -0400
Subject: [Microscopy] Re: viaWWW: video capture card for Gatan Camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I've done it, and it certainly worked first time, no problems. I used old
ISA cards, (well - they weren't old when I used them!!) I think an M-vision
MV100 and a Synoptics card - possibly a Syntillate, but I'm not sure about
that. Of course you would use newer ones, but I would be quite confident
that any one with a standard TV input would work fine. It would all depend
on the software that came with them - the "demo" software will allow you to
capture and save images, but may not be as nice as you would like. You
probably wouldn't want to get into writing your own software, and 3rd.
party software can be expensive.

Tony.

At 09:14 PM 6/8/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

*********************************************
Anthony J. Garratt-Reed, M.A., D.Phil
MIT Room 13-1027
77 Massachusetts Avenue
Cambridge, MA 02139-4307
*********************************************
Phone: (617) 253-4622
Fax: (617) 258-6479
*********************************************



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 9 07:16:40 2005



From: Faerber Jacques :      Jacques.Faerber-at-ipcms.u-strasbg.fr
Date: Thu, 9 Jun 2005 14:11:54 +0200
Subject: [Microscopy] Re: Diffusion pump oils

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Gib

In an 1997 catalog from Pfeifer/Balzers I have the following indications :

Synthetique oil 71A :

vapor pressor 2.10-8 mbar at 20°C
good chemical resistance
Good heat and oxydation resistance
Optimale pressure domaine 10E-3 to 5E-7 mbar
Ultimate pressure : {6E-9 with LN2 trap
{4E-7 with water trap
{1E-6 with air trap (what is an air trap ????)

Recommended for metalurgie, electronic tube manufacturing, E-beam
welding, sputtering, thermal evaprators, etc. Not for particules
accelerators.

So no problem for your evaporator, but it seems to be a rough cheap oil,
and not a Santovac quality one. A bit less good than DC 704.

If you send me your fax number, I can fax you the catalog sheet, but it is
in french. It gives the specifications from 61A, 71A, DC704, AN175 and
Santovac in regard.


J. Faerber
IPCMS-GSI
(Institut de Physique et Chimie des Matériaux de Strasbourg
Groupe Surface et Interfaces)
23, rue de Loess ; BP43
67034 Strasbourg CEDEX 2
France

Tel 00 33(0)3 88 10 71 01
Fax 00 33(0)3 88 10 72 48
E-mail Jacques.Faerber-at-ipcms.u-strasbg.fr

On Wed, 8 Jun 2005, Gib Ahlstrand wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Hi Listers,
}
} I'm trying to characterize two kinds of diffusion pump oils that I found in
} my "back room":
}
} 1. Balzers-Oil 71 für Diffusionspumpen. Ch 76-946
}
}
} 2. Trafo Olie, Dialac C #132250466801)
}
}
} They both look like they have been around for some years. Does anyone know
} what kind of oils they are, eg. Silicone based or not?
}
} My TEM and SEM have dedicated diff pump oils supplied by the manufacturer,
} but I'm wondering if either of the ones mentioned above would be OK to use
} in the diff pump of my Denton vacuum evaporator, which I use for routine
} carbon or gold-palladium coatings.
}
} Thanks in advance!
}
} Gib
} --
} Gib Ahlstrand, Scientist
} Electron Optical Facility, University of Minnesota, CBS Imaging Center,
} 35 Snyder Hall, St. Paul, MN. USA. 55108 (612)624-3454
} (612)624-2785 FAX, ahlst007-at-tc.umn.edu
} http://www.cbs.umn.edu/ic/
}
}
}
}




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 08:28:26 2005



From: Stacey.Andringa-at-uc.edu (by way of MicroscopyListserver)
Date: Fri, 10 Jun 2005 08:27:42 -0500
Subject: [Microscopy] viaWWW: labeling with HRP-gold and SPA-gold

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (Stacey.Andringa-at-uc.edu) from
http://www.microscopy.com/MLFormMail.html on Friday, June 10, 2005 at
08:25:18
---------------------------------------------------------------------------

Email: Stacey.Andringa-at-uc.edu
Name: Stacey Andringa

Organization: University of Cincinnati

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I have typeII cells that I want to label with HRP-gold and
SPA-gold. Someone else has made the labels for me. I have a protocol
to label the cells, but I want to make cytospin slides and check them
to see if the cells are really labelled. I have a Vector kit that
uses DAB (SK-4100).
How can I block endogenous proteins?
Thanks.

Stacey Andringa

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 12:58:47 2005



From: Brian Matsumoto :      matsumot-at-lifesci.ucsb.edu
Date: Fri, 10 Jun 2005 11:01:23 -0700
Subject: [Microscopy] Course Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


The Department of Molecular, Cellular, and Developmental Biology and the
Neuroscience Research Institute are sponsoring an advance course on light
microscopy. This 5-day workshop will be offered from September 5 through
September 9, 2005 and will consist of lectures and laboratory exercises that
will run from 9 am to approximately 5 pm each day. The seminar/workshop will
be intensive, enabling participants to develop theoretical and hands-on
expertise with light microscopes. Attendees will interact closely with the
instructors while using modern research grade microscopes, cameras, and
computers. The seminars and laboratories will cover basic optical theory and
how it pertains to increasing contrast (signal to noise ratio) in biological
samples. Fundamental techniques such as fluorescence, phase contrast,
Nomarski Differential Interference Contrast, and darkfield imaging will be
taught and attendees will use microscopes equipped with these optical
enhancement accessories. In addition, the theory and practice of electronic
image acquisition (analog and digital) will be presented and attendees will
work with low-light cameras, digital image processing computers, and
morphometric programs. There are five research grade microscopes, five
electronic imaging cameras, two computer workstations, and one confocal
microscope. With a maximum enrollment of 10 students, there will be ample
opportunity to work with all of the microscopes and cameras. For those so
interested, intensive hands-on instruction and guidance on the confocal
microscope will be provided.

For a fuller description of the workshop please check the web address below.
Enrollment forms can be completed online and this workshop provides an
opportunity to have a working-vacation in Santa Barbara, California.

http://www.lifesci.ucsb.edu/mcdb/events/imaging_workshop/index.html

Brian Matsumoto, Ph.D.
Neuroscience Research Institute and
Department of Biology: MCDB
University of California
Santa Barbara, CA 93106-9610

Phone: 805-893-8702
FAX: 805-893-4724




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 13:42:11 2005



From: Pat Connelly :      psconnel-at-sas.upenn.edu
Date: Fri, 10 Jun 2005 14:24:54 -0400
Subject: [Microscopy] Re: Kodak EM4489 films sensitivity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

6/1/04, Tindall, Randy D. wrote:
} ------------------------------------------------------------------------------
} Has as anybody noticed a second round of changes in this film? ...
}
} This leads me to think that there's a possibility that this film was
} reformulated yet again at some point, since nothing obvious changed in
} our procedures. Any similar experiences out there?
Randy,
I was justs cleaning out a lot of email messages from my file and
re-read your much longer message.
Did you ever find out if the 4489 was changed a second time?

I still have 3 boxes of 2004-06 which from my calculation should be
"original film"
but the other stock I have is 2005-06 which would fall into the time
frame that your
film batch might have been when you wrote the message. Does this make sense?

Storm just hit Bye-
Pat


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 14:03:19 2005



From: Webster, Paul :      PWebster-at-hei.org
Date: Fri, 10 Jun 2005 11:59:50 -0700
Subject: [Microscopy] Re: Re: Kodak EM4489 films sensitivity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Randy,

I didn't see your original message so my reply is late. Yes, I have noticed
a difference in the "new" 4489 film since it was changed. We have had to
reset out TEM to accommodate for the much darker images that suddenly
started to appear.

It seems like they are working at getting everyone to convert to digital in
a very covert way.

Paul.


On 6/10/05 11:24 AM, "Pat Connelly" {psconnel-at-sas.upenn.edu} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} 6/1/04, Tindall, Randy D. wrote:
} }
-----------------------------------------------------------------------------} }
-
} } Has as anybody noticed a second round of changes in this film? ...
} }
} } This leads me to think that there's a possibility that this film was
} } reformulated yet again at some point, since nothing obvious changed in
} } our procedures. Any similar experiences out there?
} Randy,
} I was justs cleaning out a lot of email messages from my file and
} re-read your much longer message.
} Did you ever find out if the 4489 was changed a second time?
}
} I still have 3 boxes of 2004-06 which from my calculation should be
} "original film"
} but the other stock I have is 2005-06 which would fall into the time
} frame that your
} film batch might have been when you wrote the message. Does this make sense?
}
} Storm just hit Bye-
} Pat
}



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 14:59:10 2005



From: Marc Pypaert :      marc.pypaert-at-yale.edu
Date: Fri, 10 Jun 2005 15:58:27 -0400
Subject: [Microscopy] Re: Re: Re: Kodak EM4489 films sensitivity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

That explains a lot!
We've been having darker negatives later too,
but I had assumed that it was either a change in
temperature or developer dilution that had caused
it.

Marc

On Friday, June 10, 2005, at 02:59 PM, Webster, Paul wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Dear Randy,
}
} I didn't see your original message so my reply is late. Yes, I have
} noticed
} a difference in the "new" 4489 film since it was changed. We have had
} to
} reset out TEM to accommodate for the much darker images that suddenly
} started to appear.
}
} It seems like they are working at getting everyone to convert to
} digital in
} a very covert way.
}
} Paul.
}
}
} On 6/10/05 11:24 AM, "Pat Connelly" {psconnel-at-sas.upenn.edu} wrote:
}
} }
} }
} } ----------------------------------------------------------------------
} } --------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} } America
} } To Subscribe/Unsubscribe --
} } http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
} -----------------------------------------------------------------------
} -------}
} -
} }
} } 6/1/04, Tindall, Randy D. wrote:
} } }
} -----------------------------------------------------------------------
} ------} }
} -
} } } Has as anybody noticed a second round of changes in this film? ...
} } }
} } } This leads me to think that there's a possibility that this film was
} } } reformulated yet again at some point, since nothing obvious changed
} } } in
} } } our procedures. Any similar experiences out there?
} } Randy,
} } I was justs cleaning out a lot of email messages from my file and
} } re-read your much longer message.
} } Did you ever find out if the 4489 was changed a second time?
} }
} } I still have 3 boxes of 2004-06 which from my calculation should be
} } "original film"
} } but the other stock I have is 2005-06 which would fall into the time
} } frame that your
} } film batch might have been when you wrote the message. Does this make
} } sense?
} }
} } Storm just hit Bye-
} } Pat
} }
}
}
}

--
Marc Pypaert
Department of Cell Biology
Center for Cell and Molecular Imaging
Ludwig Institute for Cancer Research
Yale University School of Medicine
333 Cedar Street, PO Box 208002
New Haven, CT 06520-8002
TEL 203-785 3681
FAX 203-785 7446



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 15:37:13 2005



From: Gary Gaugler :      gary-at-gaugler.com
Date: Fri, 10 Jun 2005 13:22:54 -0700
Subject: [Microscopy] [MICROSCOPY] Re: specimen heating up

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Thanks to the help from David Frey of Zeiss NY,
the problem is software related. Versions of the
SmartSEM prior to 5.00.08 had the problem. Field
service here has 09. So this should fix the problem.
If you have the heating problem, contact your service
folks for a newer software version.

PS: disabling stage R does not remove drive from the
motor. It just denies R commands.

gary g.


At 08:24 PM 6/6/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 16:22:33 2005



From: Heather A Owen :      owenha-at-csd.uwm.edu
Date: Fri, 10 Jun 2005 16:21:48 -0500 (CDT)
Subject: [Microscopy] Re: Re: Re: Re: Kodak EM4489 films sensitivity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi All,

We've been having the same problem. Kodak has also changed the
instructions on how to mix D-19 developer, though, and that may be
compounding the problem. The old (paper envelopes) instructed adding the
package to 3.8 liters of water, while the newer (plastic) envelopes have
you start with 3 liters, dissolve the solution, then bring the final
volume to 3.8 liters. Unfortunately, I threw out my last paper envelope a
few weeks before mixing a batch with the stuff in the plastic envelope so
I'm not sure whether the weight of the old (???) and new packages (551
grams) are the same or not.

Does anyone know how much powder is in a 1 gallon (3.8 liter) size paper
envelope?

Heather Owen


Dr. Heather A. Owen, Director
Electron Microscope Laboratory
Department of Biological Sciences
University of Wisconsin - Milwaukee
Lapham Hall, P.O. Box 413
Milwaukee, WI 53210
USA

Phone: (414)229-6816



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 16:27:45 2005



From: Heather A Owen :      owenha-at-csd.uwm.edu
Date: Fri, 10 Jun 2005 16:27:03 -0500 (CDT)
Subject: [Microscopy] Film Sensitivity - error in previous posting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Oops. I had an empty Dektol envelope instead of an empty D-19 envelope.
The actual weight of the plastic 3.8 liter size package is 607 grams, not
551 grams.

Friday afternoon brain slippage.

Heather Owen


Dr. Heather A. Owen, Director
Electron Microscope Laboratory
Department of Biological Sciences
University of Wisconsin - Milwaukee
Lapham Hall, P.O. Box 413
Milwaukee, WI 53210
USA

Phone: (414)229-6816



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 16:50:20 2005



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Fri, 10 Jun 2005 16:49:36 -0500
Subject: [Microscopy] Re: Re: Re: Re: Re: Kodak EM4489 films sensitivity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

heather, marc, etc

having some D-19 of older stock (just getting into the next order), the
package size with the paper bag is 12gms less than for the new plastic
bags. this difference could be just packaging. oh, yeah, that is for
D-19, not dektol. our friday afternoon brain seizure has not yet hit up
her in the frozen north....

as far as technique, the original procedure in my books - like 30+ years
ago - was make it up in 3L at 37-50C, let cool, and then bring up to
3.8L volume. that is the way i've always done it wherever i have gone.

hope that trivia helps.

paul



Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-9354
Cell:204-781-1502
Fax:204-789-3926



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 10 16:56:29 2005



From: Greg Strout :      gstrout-at-ou.edu
Date: Fri, 10 Jun 2005 16:55:45 -0500
Subject: [Microscopy] Re: Re: Re: Re: Re: Kodak EM4489 films sensitivity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Heather,
I don't know about the paper bags as we don't have any of those, but we
do have the cans that pre-date the paper bags and they have 595 grams
with the instructions to add powder to 1 U.S. gallon (3.8 liters) at
100F (38C).
Greg

Heather A Owen wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
--
==================================================================
Greg Strout
Electron Microscopist, University of Oklahoma
WWW Virtual Library for Microscopy:
http://www.ou.edu/research/electron/www-vl/
e-mail: gstrout-at-ou.edu
Opinions expressed herein are mine and not necessarily those of
the University of Oklahoma
==================================================================




From MicroscopyL-request-at-ns.microscopy.com Sat Jun 11 22:14:04 2005



From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sat, 11 Jun 2005 23:13:23 -0500
Subject: [Microscopy] SEM/EDS service labs in China

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Stu Smalinskas wrote:
==============================================
I just got a request from someone in China if there
are any labs available in China that can do SEM/EDS
work. Can anyone provide contact information?
==============================================
See URL
http://www.2spi.com/agntdist/china.html

The first firm listed is run by Dr. Susan Tai, formerly an executive with
FEI Far East. She knows everyone who does this kind of work in China.

The third firm, KYKY, is a Chinese manufacturer of SEMs and at least at one
time would run commercial samples for clients in their demo lab in Beijing.
I have seen their newest model operate and it produces quite a good image.
They are physically located next door to the large and very well equipped
lab of the Academia Sinica (Chinese Academy of Sciences) so if they could
not do what was needed, I am sure the lab next door could.



Chuck

===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: http://www.2spi.com
############################
==================================================




From MicroscopyL-request-at-ns.microscopy.com Sat Jun 11 23:07:13 2005



From: Ellery Frahm :      frah0010-at-tc.umn.edu
Date: Sat, 11 Jun 2005 23:06:31 -0500
Subject: [Microscopy] lab noise control and acoustic panels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy folks,

Our electron microprobe laboratory has a chiller and sizable air
conditioner, and thanks to Facilities Management at our university,
there's little chance of getting them moved out of the lab. Since I
cannot move the noise sources out of the lab, I've been thinking
about installing acoustic panels to help with noise control. I've
seen this done at a few other electron microscope labs, and the staff
seemed happy with the results. I've been looking online at acoustic
panels, and there are various materials, shapes, and designs out
there. Has anyone else done this? What materials and shape did you
use? Are you happy with the results? Any problems (shedding of
fibers or dust, etc)? Is there anything you wish you'd done
differently? Any suggestions or opinions are welcome.

Thanks in advance,

Ellery

--------------------
Ellery E. Frahm
Research Fellow & Manager
Electron Microprobe Laboratory
University of Minnesota - Twin Cities
Department of Geology & Geophysics
Lab Website: http://probelab.geo.umn.edu/



From MicroscopyL-request-at-ns.microscopy.com Sun Jun 12 11:08:16 2005



From: Larry Allard :      allardlfjr-at-ornl.gov
Date: Sun, 12 Jun 2005 12:07:01 -0400
Subject: [Microscopy] Re: lab noise control and acoustic panels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ellery:

I personally would avoid the use of foam-based materials such as
Sonex, as they degrade with time (~10years) and become a real mess to
clean up and replace. We have begun using absorber/barrier blanket
material, such as provided by Netwell Noise Control (and others),
which does a great job and looks like it will last virtually forever.
Get on Google and check it out. You can have special enclosures made
to your own design, to isolate pumps and chillers, for example.
Larry





} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
Dr. Lawrence F. Allard
Distinguished Research Staff Member
High Temperature Materials Laboratory
Microscopy, Microanalysis, Microstructures Group
Metals and Ceramics Division
Oak Ridge National Laboratory
1 Bethel Valley Road
PO Box 2008
Oak Ridge, TN 37831-6064
(note: the last 4 lines are sufficient for mailing or overnight
courier service)
865-574-4981
865-576-5413 Fax
http://www.ms.ornl.gov/htmlhome/mauc


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 13 00:43:33 2005



From: Lorina Dillon :      qfzaluzec-at-microscopy.com
Date: Mon, 13 Jun 2005 01:42:45 -0500
Subject: [Microscopy] enhance ur S,exual performance up to 24-36

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Read This

Ci-alis Softtabs

Get ready for romance in just 15minutes
Increase your Se,x Drive
Boost Se'xual Performance
Fuller and Harder Er-ections
Increase Stamina and Endurance
Quicker Recharges
New Super Via-gra (100x better)
Discreet online Orde,ring and shipping
DHL Express Delivery

Special Ends Soon... 0rder Here -} http://YYcdwW.impactions.net/cs/?unxx

Bye

Lorina Dillon


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 13 06:51:22 2005



From: Michael Cheatham :      mmcheath-at-mailbox.syr.edu
Date: Mon, 13 Jun 2005 07:48:59 -0400
Subject: [Microscopy] Re: lab noise control and acoustic panels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ellery,
I haven't done this with our SEM lab, but I have used sound
baffles in our Noble Gas lab. That lab has two mass specs with 7
rough pumps and 5 turbo pumps between them. Those pumps plus all of
the electronic unit cooling fans add a fair amount of noise to the
lab. I originally wanted sound baffling mounted to the walls, but
our design and construction people would not allow it. Instead they
recommended hanging 2' x 4' sound baffles from the suspended ceiling
grid. They gave me a recommended layout which I tried to follow as
best I could. The overall effect was excellent. The baffles we used
are basically a tight weave of fiberglass enclosed in a poly sealed
'bag'. No dust collects on them as happens with wall mount egg crate
material. They had two grommet wholes along one edge and they came
with standard T-bar clips to hang them. I just used zip ties to
connect the grommet to the hangers. Unfortunately, I don't know who
makes the products as our design people did all the purchasing and
had everything delivered to the lab. I hung them myself because I
don't trust our facilities people working above our instruments.

Mike


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
********************************************************************
Michael M. Cheatham
312 Heroy Geology Laboratory Phone (315)-443-1261
Syracuse University Fax (315)-443-3363
Syracuse, NY 13244-1070

email:mmcheath-at-mailbox.syr.edu
http://www.geochemistry.syr.edu

owner of PLASMACHEM-L: http://listserv.syr.edu/archives/plasmachem-l.html
owner of XRF-L: http://listserv.syr.edu/archives/xrf-l.html
owner of TIMS-L: http://listserv.syr.edu/archives/tims-l.html
owner of SIRIS-L: http://listserv.syr.edu/archives/siris-l.html
********************************************************************


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 13 09:37:14 2005



From: Debby Sherman :      dsherman-at-purdue.edu
Date: Mon, 13 Jun 2005 09:10:33 -0500
Subject: [Microscopy] Re: M&M2005 session and noise control

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Ellery and others,

I have also run into some decision making regarding acoustic panels for a
high end FESEM room. This is of major concern in a growing number of labs
installing the newer very high end TEMs and SEMs. We are going to have a
platform presentation about designing a high end facility in the Core
Facility Management session at M&M2005. The session is from 1:30 to 3:30 on
Wednesday.

I would really appreciate those who have information on acoustic noise
remedies (as well as vibration and electrical interference remedies) to
contact me. If you are going to be at M&M2005 I would invite you to
participate in this session to share this information.

Another topic that will be introduced during the session, with additional
discussion to follow the session (either at the Convention Center or at
dinner Wednesday evening) is that of LIM (large Image Management) software
and related issues (large data storage and backup, billing, authorization).

The discussion will be introduced by Avrum Goodblat and will revolve around
two major issues:

a. LIMS software
b. how to hook LIMS to the microscope. This is not always so easy to do - it
depends both on the software and the nature of the utilities provided by the
microscope manufacturer.

The Core Facility Management session, organized by the Focused Interest
Group on Facility Operation, is designed to address topics of current
interest. These threads often appear on the list serve a few months before
the actual meeting and cannot be included in the meeting program. Since the
session is geared to discussion of current topics of interest, we have a
great deal of flexibility. Please take advantage of this and plan to bring
information and participate fully in these discussions.


On 6/11/05 11:06 PM, "Ellery Frahm" {frah0010-at-tc.umn.edu} wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------------}
-
}
} Microscopy folks,
}
} Our electron microprobe laboratory has a chiller and sizable air
} conditioner, and thanks to Facilities Management at our university,
} there's little chance of getting them moved out of the lab. Since I
} cannot move the noise sources out of the lab, I've been thinking
} about installing acoustic panels to help with noise control. I've
} seen this done at a few other electron microscope labs, and the staff
} seemed happy with the results. I've been looking online at acoustic
} panels, and there are various materials, shapes, and designs out
} there. Has anyone else done this? What materials and shape did you
} use? Are you happy with the results? Any problems (shedding of
} fibers or dust, etc)? Is there anything you wish you'd done
} differently? Any suggestions or opinions are welcome.
}
} Thanks in advance,
}
} Ellery
}
} --------------------
} Ellery E. Frahm
} Research Fellow & Manager
} Electron Microprobe Laboratory
} University of Minnesota - Twin Cities
} Department of Geology & Geophysics
} Lab Website: http://probelab.geo.umn.edu/
}
}




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 13 12:15:52 2005



From: micrograph05-at-aol.com
Date: Mon, 13 Jun 2005 13:15:02 -0400
Subject: [Microscopy] LM Small World Competition - Final Call for Entries

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


www.nikonsmallworld.com

Attention Microscopists,

June 30, 2005 is the final deadline for submitting entries to the 31st
Nikon International Small World Competition. The competition, open to
all microscopists, was created in 1975 to recognize excellence in
photography through the light microscope, and to applaud the efforts of
those involved with photomicrography.

The competition is open to all types of light microscopy, and there are
no restrictions to the type of specimens or techniques used
(brightfield, darkfield, phase contrast, interference contrast,
fluorescence, confocal, deconvolution, mixed techniques, etc.).
Participants may obtain the rules, entry forms and enter their digital
images directly at http://www.nikonsmallworld.com/. For those
submitting entries in film format, the image and entry form should be
sent to: Nikon Small World, 1300 Walt Whitman Road, Melville, NY
11747-3064. The first and second of twenty prizewinners will receive a
selection of Nikon products and equipment worth $3,000 and $2,000
respectively.

Winners will be selected mid-summer. Each year, exhibits containing the
winning entries are displayed at museums and science centers throughout
North America. Many of these spectacular images have been featured on
the covers of prestigious scientific publications and journals.


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 13 13:56:22 2005



From: biology :      biology-at-ucla.edu
Date: Mon, 13 Jun 2005 11:55:37 -0700
Subject: [Microscopy] EM Position Opening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



http://jobs.healthcare.ucla.edu/03_header_job.html

Job Title: Histotechnologist III/EM
UCLA Title: Histotechnologist III
Job No.: H33530
Work Hours: Monday-Friday, 8:00am-5:00pm
Work Location: CHS A3-240
Minimum Salary: $20.98 / $3651
Maximum Salary: $29.37 / $5110
Layoff Referral Deadline:

Job Duties:
Provide administrative, technical, and maintenance support to the Electron
Microscopy Department. Process, embed and cut EM blocks and tissue while
maintaing excellent quality. Prepare: thin sections, staining, special
processing, preparation of knives, reviewing and scanning of grids, and
digital photography. Peforms independently a wide variety of technical and
non-technical procedures. Perform processing of primary specimens, reagent
preparation, operating, cleaning and maintaing work area and laboratory
equipment, troubleshooting specimen/test order problems, filing, answering
phones to respond to questions and participating in special
studies. Oversee staff in their assigned area to see that all procedures
meet the protocol stated. Tract laboratory supplies and places orders as
necessary. Monitoring inventories of media, supplies and reagents used in
diagnostic testing, quality control/quality assurance support, computer
oepration (report printing) and data entry and retrieval. Enter data into
the Tamtron system, maintain and print logs and labels. Perform daily
recordkeeping.


Job Qualifications:
Working knowledge of basic chemistry, biology, and anatomy with minimum of
5 years experience in Electron Microscopy. Skill in sectioning difficult
tissues. Skill in performing all ranges of testing and procedures in
Electron Microscopy. Ability to learn running variety of automated
stainers and other equipments. Skill in Tamtron system or ability to learn
very fast. Ability to maintain the work area clean and follow the safety
procedure. Ability to speak clearly and distinctly using appropriate
English vocabulary. Ability to communicate effectively with resident and
faculty in regard to the different issues. Ability to communicate with
mutual respect and consideration for diversity among individual. Ability
to work as part of team and cover the missing area voluntarily. Able to
maintain confidentiality. Willingness to participate in teleconferences
and other continuing education program. Able to modify work schedule to
meet the needs of the department. Ability to maintain the quality
control. Ability to perform any other duties or tasks as requested by the
sueprvisor.





From MicroscopyL-request-at-ns.microscopy.com Mon Jun 13 16:26:37 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 13 Jun 2005 14:25:25 -0700
Subject: [Microscopy] F30H objective aperture & low dose

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Lists,
A problem has come up with using the objective aperture in conjunction
with the low dose kit. When the aperture is centered in the Exposure
state, it is off-center in the Focus state by an amount that varies
with the focus shift. The problem is the same using TEM mode or EFTEM
mode. Logically, this should be related to image shift pivot points,
but that alignment is right on, and a complete set of TEM-HM alignments
did not solve the problem. Both our local service guy and an expert
FEI guy have looked at the problem, but have not been able to solve it
(although we thought they had on a couple of occasions). Has anyone
else seen this, and, if so, what fixed it. TIA.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 13 20:11:25 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 13 Jun 2005 18:10:15 -0700
Subject: [Microscopy] Re: [3DEM] F30H objective aperture & low dose

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Jun 13, 2005, at 2:57 PM, Jamie Riches wrote:

} I was of the understanding that
} this problem was a fundamental one related to the image shifts and the
} objective apertures, but if you find out otherwise, could you please
} send an
} email to either of the lists?
}
Dear Lists,
According to David Mastronarde, the problem is inherent in the Tecnai,
since the obj aper is not in the back focal plane, and the alignment
procedures do not account for this.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 08:34:22 2005



From: Barbara Maloney :      maloneyb-at-fiu.edu
Date: Tue, 14 Jun 2005 09:30:36 -0400
Subject: [Microscopy] Current usage of TEM or FTEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Group - what do you believe current usage for TEM or field emission
TEM is right now for biological applications?
Thanks in advance for your input
Barbara


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 09:13:37 2005



From: Sherwood, Margaret :      MSHERWOOD-at-PARTNERS.ORG
Date: Tue, 14 Jun 2005 10:12:54 -0400
Subject: [Microscopy] Current usage of TEM or FTEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Barbara,et al:

It's funny. For the first 16 years I have been at Wellman, we had two full-time
TEM people (myself included) who had non-stop projects to do. Then, about 3
years ago, (somewhat coinciding with the purchase of the confocal
microscope),TEM work dried up; literally. Now, with only myself running it, TEM
work is picking up again. I currently have 3 projects to scope. Our work is
stricly biological (I am looking at bacteria, retinal lesions and tattooed
skin). We are a large laser research center at MGH and we collaborate with a
number of the departments. (I don't just do the TEM for the group; I train
people on the confocal and am doing molecular biology as well).

We are still not up to the volume of work (TEM) that we did and probably will
never reach that level of work again, but it is nice to get back to basic
ultrastructure.

Peggy

Peggy Sherwood
Lab Associate, Photopathology
Wellman Center for Photomedicine (W224)
Massachusetts General Hospital
55 Fruit Street
Boston, MA 02114
617-724-4839 (voice mail)
617-726-6983 (lab)
617-726-3192 (fax)
msherwood-at-partners.org



-----Original Message-----
} From: Barbara Maloney [mailto:maloneyb-at-fiu.edu]
Sent: Tuesday, June 14, 2005 9:31 AM
To: microscopy-at-msa.microscopy.com

Dear Group - what do you believe current usage for TEM or field emission
TEM is right now for biological applications?
Thanks in advance for your input
Barbara




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 11:38:13 2005



From: Pat Connelly :      psconnel-at-sas.upenn.edu
Date: Tue, 14 Jun 2005 12:17:00 -0400
Subject: [Microscopy] Re: Current usage of TEM or FTEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} Dear Group - what do you believe current usage for TEM or field emission TEM
} is right now for biological applications?
} Thanks in advance for your input
} Barbara {maloneyb-at-fiu.edu}
==========================
Barbara,
I am currently the only TEM Technologist in my biology department and
since my PI is retiring in September it will also be the end of my
employment here. Over the past several years we have had some
collaborators who could not get research quality TEM done in their
own departments or institutions. They realized that biochemistry,
confocal microscopy, etc. showed that the experiments were working
but could not give the ultrastructural confirmation that was
ultimately needed for understanding and hence to get the publication
of their data accepted. More than once I have heard the words, "That
is exactly what I thought was happening but I could not prove it."

A grant request has been made for a new TEM for our department, the
present scope being 39 years old. Ten investigators are making plans
to use the scope (hoping that the grant is funded) so it seems that
there is still a need for TEM [and I might add that I am attempting
to make known to them that it may be wise to have an experienced
technologist around!].

For these reasons, I state that TEM has not been abandoned only "put
on a back burner" while other techniques have caught the glamor of
the day. As more investigators see the importance of confirming
their results on the ultrastructural level or re-discover the sense
of wonder at seeing what has not been observed or reported
previously, TEM will regain its place in biological applications.

Just my thoughts and feelings,
Pat Connelly psconnel-at-sas.upenn.edu
Dept. of Biology
Univ. of Pennsylvania
Philadelphia, PA






From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 12:36:54 2005



From: :      nyilmaz-at-mersin.edu.tr
Date: Tue, 14 Jun 2005 20:36:09 +0300
Subject: [Microscopy] staphylococcus suspension protocol

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello everybody...

I'm searching a protocol processing staphylococcus suspension for TEM.

Thanks in advance...

Dr. Nejat Yilmaz







From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 13:20:43 2005



From: Leona Cohen-Gould :      lcgould-at-med.cornell.edu
Date: Tue, 14 Jun 2005 14:19:55 -0400
Subject: [Microscopy] Re:mouse embryo SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi All,
I was just asked about how to process mousse embryos (embryonic days
9.5, 10.5 and 11.5) for SEM specifically to look at the limb buds.
I"ve done TEM of limb buds, but not SEM. Would anyone out there care
to suggest a protocol (if it differs substantially from "standard"
fixation-dehydration and CPD)?
Thanks,
Lee
--
Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy & Histology Core Facility
Manager, Optical Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 14:25:21 2005



From: Frank Macaluso :      macaluso-at-aecom.yu.edu
Date: Tue, 14 Jun 2005 15:24:35 -0400
Subject: [Microscopy] Re: Re:mouse embryo SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Lee,
When we processed mouse embryos for SEM ( in '92) we had excellent results
using "standard" fixation, dehydration and CPD.
Frank

At 02:19 PM 6/14/2005, Leona Cohen-Gould wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 14:55:19 2005



From: Rhonda Stroud :      stroud-at-nrl.navy.mil
Date: Tue, 14 Jun 2005 15:54:26 -0400
Subject: [Microscopy] TEM/FIB postdoc position available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



The Materials Science Division at the Naval Research Laboratory in
Washington, DC, seeks postdoctoral fellow candidates to carry out TEM and
FIB-based investigations of nanostructured materials. Potential projects
include analysis of natural and synthetic nanoparticles, e.g., 1 to 10 nm
particles condensed in the upper atmosphere and/or laboratory grown by
reverse micelle synthesis; FIB-based micromanipulation and analysis of
cosmic dust samples (STARDUST Mission and dust from stars pre-dating the
Sun), and in situ STM-TEM studies of nanowires. NRL has extensive
facilities for microscopy studies, including newly installed JEOL 2200FS
and FEI Nova 600 DB-FIB instruments. Experience in HAADF imaging, EDS and
EELS spectrum imaging and/or FIB operation are desired. A Ph.D. in
physics, chemistry, materials science, or geology is required. Qualified
candidates will be expected to submit a proposal for the August 1, NRC
fellowship deadline, but start dates prior to the October completion of the
review process are possible. US citizens or permanent residents only. For
more information, contact Rhonda Stroud (stroud-at-nrl.navy.mil).




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 15:33:00 2005



From: David Elliott :      elliott-at-arizona.edu
Date: Tue, 14 Jun 2005 13:32:12 -0700
Subject: [Microscopy] Re: Re:mouse embryo SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

http://www3.interscience.wiley.com/cgi-bin/abstract/102524606/ABSTRACT
Has the SEM embedding procedures

Grose R, Harris BS, Cooper L, Topilko P, Martin P. 2002. Immediate
early genes krox-24 and krox-20 are rapidly up-regulated after
wounding in the embryonic and adult mouse. Dev Dyn 223: 371-378. Links
Has the exact fix

David



On Jun 14, 2005, at 11:19 AM, Leona Cohen-Gould wrote:

}
}
} ----------------------------------------------------------------------
} --------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/
} MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} ----------------------------------------------------------------------
} ---------
}
} Hi All,
} I was just asked about how to process mousse embryos (embryonic
} days 9.5, 10.5 and 11.5) for SEM specifically to look at the limb
} buds.
} I"ve done TEM of limb buds, but not SEM. Would anyone out there
} care to suggest a protocol (if it differs substantially from
} "standard" fixation-dehydration and CPD)?
} Thanks,
} Lee
} --
} Leona Cohen-Gould, M.S.
} Sr. Staff Associate
} Director, Electron Microscopy & Histology Core Facility
} Manager, Optical Microscopy Core Facility
} Joan & Sanford I. Weill Medical College
} of Cornell University
} voice (212)746-6146
} fax (212)746-8175
}
}



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 15:35:55 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Tue, 14 Jun 2005 13:34:46 -0700
Subject: [Microscopy] Re: Current usage of TEM or FTEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Jun 14, 2005, at 6:30 AM, Barbara Maloney wrote:

} what do you believe current usage for TEM or field emission TEM is
} right now for biological applications?
} Thanks in advance for your input

Dear Barbara,
All structural biology benefits from TEM (although not all uses it).
We are doing both tomography, for high-resolution 3-D imaging of
biological specimens, and single-particle analysis, for
very-high-resolution determination of the structures of protein
complexes. Almost all our work in on frozen-hydrated unstained
specimens, which is as close to the native state as can presently be
achieved. So the bottom line is that (FE)TEM is the best technique for
determining the high-resolution structure of (nearly) native biological
specimens.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 14 15:39:49 2005



From: Larry Stoter :      larry-at-cymru.freewire.co.uk
Date: Tue, 14 Jun 2005 21:35:41 +0100
Subject: [Microscopy] Re: lab noise control and acoustic panels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Curtains? Works well in many cases and a lot cheaper than specialist
panels. If they don't work, you won't have lost much; if they do,
you'll probably have saved quite a bit.
--
Larry Stoter
JEOL (UK) Ltd
tel: +44-(0)1707-377117, fax: +44-(0)1707-373254, e-mail: larrys-at-jeoleuro.com

PLEASE NOTE
1. Any mail other than plain text will be automatically deleted.
2. Any mail, legitimate or not, apparently or actually from hotmail,
netscape, yahoo or excite will automatically be deleted.
3. Mail with no subject or without a clear subject will be ignored :-)


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 15 11:39:16 2005



From: Dorota Wadowska :      wadowska-at-upei.ca
Date: Wed, 15 Jun 2005 13:13:57 -0400
Subject: [Microscopy] LM fish eggs processing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Everybody,
I have a user who wants to process fish eggs (salmonids) for LM.
He embedded eggs in paraffin but infiltration was insufficient and
samples could not be cut at all. Is there a better protocol to handling
these kind of samples? We tried Spurr with dehydration time
extended to 1h in each concentration of alcohol and resin infiltration
overnight. Still did not work. The interior was not properly infiltrated
and sample was shrunk.
Thanks
Dorota


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 15 15:07:20 2005



From: Larry Stoter :      larry-at-cymru.freewire.co.uk
Date: Wed, 15 Jun 2005 20:48:37 +0100
Subject: [Microscopy] Re: Re: lab noise control and acoustic panels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

} On 6/14/05 3:35 PM, "Larry Stoter" {larry-at-cymru.freewire.co.uk} wrote:
}
} }
} ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
} ------------------------------------------------------------------------------}
} -
} }
} } }
} -----------------------------------------------------------------------------} }
} -
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} } } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } }
} } } -----------------------------------------------------------------------------
} } } --
} } }
} } } Microscopy folks,
} } }
} } } Our electron microprobe laboratory has a chiller and sizable air
} } } conditioner, and thanks to Facilities Management at our university,
} } } there's little chance of getting them moved out of the lab. Since I
} } } cannot move the noise sources out of the lab, I've been thinking
} } } about installing acoustic panels to help with noise control. I've
} } } seen this done at a few other electron microscope labs, and the
} } } staff seemed happy with the results. I've been looking online at
} } } acoustic panels, and there are various materials, shapes, and
} } } designs out there. Has anyone else done this? What materials and
} } } shape did you use? Are you happy with the results? Any problems
} } } (shedding of fibers or dust, etc)? Is there anything you wish you'd
} } } done differently? Any suggestions or opinions are welcome.
} } }
} } } Thanks in advance,
} } }
} } } Ellery
} } }
} } } --------------------
} } } Ellery E. Frahm
} } } Research Fellow & Manager
} } } Electron Microprobe Laboratory
} } } University of Minnesota - Twin Cities
} } } Department of Geology & Geophysics
} } } Lab Website: http://probelab.geo.umn.edu/
} }
} } Curtains? Works well in many cases and a lot cheaper than specialist
} } panels. If they don't work, you won't have lost much; if they do,
} } you'll probably have saved quite a bit.

} Larry,
} Can you provide any specifics about curtains such as what to look for,
} possible sources, etc.
}
} Debby
}
} Debby Sherman, Manager Phone: 765-494-6666
} Life Science Microscopy Facility FAX: 765-494-5896
} Purdue University E-mail: dsherman-at-purdue.edu
} S-052 Whistler Building
} 170 S. University Street
} West Lafayette, IN 47907
} http://www3.agriculture.purdue.edu/microscopy
}

Nothing special - in the JEOL UK Applications Lab we just went to the
local department store and ordered up some curtains of the correct
size and hung them against the wall, ceiling to floor. Went for a
fairly heavy, non-synthetic material, as we thought synthetics might
cause problems with static electricity. Even to my ears, the room
acoustics were very noticeably deadened. They weren't exactly cheap,
since we needed quite a lot and it was quite a heavy material. On the
other hand, I'd bet it cost less than specialist acoustic tiles.

--
Larry Stoter
JEOL (UK) Ltd
tel: +44-(0)1707-377117, fax: +44-(0)1707-373254, e-mail: larrys-at-jeoleuro.com

PLEASE NOTE
1. Any mail other than plain text will be automatically deleted.
2. Any mail, legitimate or not, apparently or actually from hotmail,
netscape, yahoo or excite will automatically be deleted.
3. Mail with no subject or without a clear subject will be ignored :-)


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 15 20:08:38 2005



From: dgarrett-at-unt.edu (by way of MicroscopyListserver)
Date: Wed, 15 Jun 2005 20:07:54 -0500
Subject: [Microscopy] viaWWW: SEM general JEOL 5800 CMD list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (dgarrett-at-unt.edu) from http://www.microscopy.com/MLFormMail.html on Wednesday, June 15, 2005 at 10:06:28
---------------------------------------------------------------------------

Email: dgarrett-at-unt.edu
Name: Da vid Garrett

Organization: University of North Texas

Title-Subject: [Microscopy] [Filtered] SEM general JEOL 5800 CMD list

Question: Can anyone supply me with a list of the features that are controled with the CMD key. To change the load current press CMD type bias use focus and mag to change. Thats the only one I know.
Thanks, David

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 07:41:24 2005



From: chenkaic-at-msu.edu (by way of MicroscopyListserver)
Date: Thu, 16 Jun 2005 07:40:41 -0500
Subject: [Microscopy] viaWWW: EPON and Prion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (chenkaic-at-msu.edu) from http://www.microscopy.com/MLFormMail.html on Wednesday, June 15, 2005 at 23:27:20
---------------------------------------------------------------------------

Email: chenkaic-at-msu.edu
Name: Kai-Chun

Organization: MSU

Title-Subject: [Microscopy] [Filtered] EPON and Prion

Question: Dear all,

My advisor and I want to visulize prion protein with EM.
We want to use negative stainning.
However, it is somewhat difficult because of the safety level of EM lab, so we are thinking about how to deactivate prion.
I heard that EPON may deactivate prion, but I can not find the references to support that.
I wonder if anyone has this kind of experiences to handle prion protein.
Do you think EPON can deactivate prion protein?

Thank you very much!

Kai-Chun



---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 09:22:06 2005



From: paul r hazelton :      paul_hazelton-at-umanitoba.ca
Date: Thu, 16 Jun 2005 09:21:21 -0500
Subject: [Microscopy] Re: prions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

help!!!

i seem to have lost a very good response to the recent posting asking
about prion infectivity - raised by ralph common. someone on the list
responded with information from previous work of theirs, where they had
checked to determine retention of catalytic infectivity of prions after
osmication.

i seem to not have that particular response saved. could any one on the
list who has the note, or the original respondant, please send me
another copy - it was excellent, probably the best of the whole group of
responses. i wish to have it in my notes for future reference.

thanks

paul

Paul R. Hazelton, PhD
Electron Microscope Unit
University of Manitoba
Department of Medical Microbiology
531 Basic Medical Sciences Building
730 William Avenue
Winnipeg, Manitoba, Canada, R3E 0W3
e-mail: paul_hazelton-at-umanitoba.ca
Phone:204-789-3313
Pager:204-931-9354
Cell:204-781-1502
Fax:204-789-3926



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 09:27:58 2005



From: Aleksandr Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Thu, 16 Jun 2005 15:27:40 +0100
Subject: [Microscopy] Re: viaWWW: EPON and Prion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Kai-Chun,

} Question: Dear all,
}
} My advisor and I want to visulize prion protein with EM.
} We want to use negative stainning.
} However, it is somewhat difficult because of the safety level of EM lab, so we are thinking about how to deactivate prion.
} I heard that EPON may deactivate prion, but I can not find the references to support that.
} I wonder if anyone has this kind of experiences to handle prion protein.
} Do you think EPON can deactivate prion protein?
}
}
IMHO, the main danger when you work with prion protein infected tissue
is the possible inhalation or ingestion of the contaminated material.
Therefore, if you will cover the grid with some strong film then it will
in theory protect the environment from the prion exposure. However, all
the films have the tendency to break at some point that makes their
protection questionable. The idea of covering the grid with Epon may
work because Epon is quite resistant material. However in this case the
resolution can be decreased because of the thickness of epoxy film
cover. The best solution to see negatively stained prion protein on the
grid would be dedicated microscope. It is not cheap solution but it is safe.
If you will embedd your protein in epon and then make sections then
probably this protein would be inactive, but I do not have any
references that prove it.
You may want to contact Dr. Holger Wille from S.Prusiner lab (UCSF) who
has vast experience with prion proteins in EM.

Sincerely,

--
Aleksandr Mironov
Experimental Officer
G450A, Stopford Building
EM Unit, Faculty of Life Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

Tel. +44-(0)161-275-7395
Fax. +44-(0)161-275-5171
E-mail: Aleksandr.Mironov-at-manchester.ac.uk
MSN: amironov-at-hotmail.co.uk
Web: http://www.empgu.man.ac.uk




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 09:37:01 2005



From: Karl Hagglund :      hagglundk1-at-nku.edu
Date: Thu, 16 Jun 2005 10:36:19 -0400
Subject: [Microscopy] Need JEE 4B coater parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I've brought an old high vacuum coater back to life, and it is working
well for carbon coating, but I am missing the parts to do metal coatings
and set up an aperture cleaning boat.

The unit is a Jeol JEE 4B, manufactured around 1972. What I am looking
for is not all that different from ring stand parts, and I wonder
whether anyone out there has modified ring stand parts to use with a
high vacuum coater. It seems that all I would need is a couple of
insulators and a tap and die set to attach a piece of wire or moly boat
and get going.

Otherwise, is there a third party company that makes parts that would
fit on one of these?

Thanks for your help.

Karl Hagglund
Biological Sciences, SC-102B
Northern Kentucky University, Nunn Drive
Highland Heights, KY 41099
859-572-5238
http://semlab.nku.edu




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 10:29:30 2005



From: Barbieri Thomas-r53545 :      Thomas.Barbieri-at-freescale.com
Date: Thu, 16 Jun 2005 08:28:45 -0700
Subject: [Microscopy] home-built voltage contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Hello Listers,

We'd like to scratch-build an apparatus that allows us to perform voltage contrast testing in one of our SEMs. We currently have a Hitachi S-4500, a JEOL 6301, and a JEOL 6340. Does anyone on here have experience using or building such a system? If you do, could you provide a description of the components and provide hints and tips on assembling one?

Regards,
Tom

Thomas Barbieri
Arizona Product Development and Analysis Laboratory
Freescale Semiconductor
Tempe, AZ
480-413-4007

thomas.barbieri-at-freescale.com

The information contained in this email and any attachment is considered:
[X] General Business Information
[ ] Freescale Internal Use Only
[ ] Freescale Confidential Proprietary


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 10:32:53 2005



From: Bill Powell :      whpowell-at-sbcglobal.net
Date: Thu, 16 Jun 2005 12:24:57 -0400
Subject: [Microscopy] Need JEE 4B coater parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Paul & Listers:

It was me...here is the text from that thread...


Hi All:

Having started my professional career managing a EM lab working on CJD I
have actual experience working with this fun stuff. We found that PF &
Glut perfusions did not inactivate the mutant protein. Once the tissue was
post-fixed in OsO4 we could not serial pass the disease again. Once in
plastic it's for all intents and purposes inert. Waste materials were
autoclaved for a longer length of time than normal and discarded with
"normal" medical waste. In the early 80's when I was working with CJD it
was a BSL2 for the most part, now all procedures should be done under BSL3.
The Neuropathologist used to work with it under BSL1, removing the CJD
infected brains from the cadavers. They would always be violating
containment protocols in some was shape or form. They are still alive,
except one (CVA). The point I'm trying to make is don't over react, take
all of the prudent precautions of BSL3 you will be well protected.


Al Coritz
Electron Microscopy Sciences
www.emsdiasum.com

----- Original Message -----
} From: "paul r hazelton" {paul_hazelton-at-umanitoba.ca}
To: "Ralph Common" {Ralph.Common-at-hc.msu.edu}
Cc: {Microscopy-at-microscopy.com}
Sent: Thursday, June 16, 2005 10:21 AM

Karl,

Thank you for the conversation this afternoon.

Items for the JEE-4B may be available from our Parts Department (Nancy
Green). Instruction manuals (Ion Gauge) may be a bit more difficult however
we may be able to provide copies.

The main number to JEOL USA Inc in Peabody Massachusetts is 978-535-5900,
please ask for Nancy Green.

I will let you know when we have fixed the schedule for UC on the JEM-1230
demonstration.

Nice to have spoken with you.

Regards,

Bill Powell
JEOL USA Inc
Phone: 248-366-8351
Ion Beam Cross Section Polisher:
http://www.jeol.com/spe/speprods/cross_section.html


-----Original Message-----
} From: Karl Hagglund [mailto:hagglundk1-at-nku.edu]
Sent: Thursday, June 16, 2005 10:36 AM
To: microscopy-at-microscopy.com

I've brought an old high vacuum coater back to life, and it is working
well for carbon coating, but I am missing the parts to do metal coatings
and set up an aperture cleaning boat.

The unit is a Jeol JEE 4B, manufactured around 1972. What I am looking
for is not all that different from ring stand parts, and I wonder
whether anyone out there has modified ring stand parts to use with a
high vacuum coater. It seems that all I would need is a couple of
insulators and a tap and die set to attach a piece of wire or moly boat
and get going.

Otherwise, is there a third party company that makes parts that would
fit on one of these?

Thanks for your help.

Karl Hagglund
Biological Sciences, SC-102B
Northern Kentucky University, Nunn Drive
Highland Heights, KY 41099
859-572-5238
http://semlab.nku.edu






From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 12:28:25 2005



From: Evelyn York :      eyork-at-ucsd.edu
Date: Thu, 16 Jun 2005 10:27:43 -0700
Subject: [Microscopy] Re: home-built voltage contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Tom,

There are more then a few variables to consider when you decide to utilize
voltage contrast. For example, in addition to the tooling and fixtures and
SEM, the device type and fabrication process play a huge role in how
successful you can be at capturing the correct electron signals. Please
contact me off line and I'll share whatever experience and knowledge I have.

Regards,
Evelyn


At 08:28 AM 6/16/2005, Barbieri Thomas-r53545 wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Evelyn York

Analytical Facility
Scripps Institution of Oceanography
University of California, San Diego
9500 Gilman Drive M/S 208
La Jolla, CA 92093-0208

(858) 534-2438



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 16 21:03:37 2005



From: emlab-at-vet.ksu.edu (by way of MicroscopyListserver)
Date: Thu, 16 Jun 2005 21:02:54 -0500
Subject: [Microscopy] viaWWW: transferrine

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (emlab-at-vet.ksu.edu) from http://www.microscopy.org/MicroscopyListserver/MLFormMail.html on Thursday, June 16, 2005 at 08:51:27
---------------------------------------------------------------------------

Email: emlab-at-vet.ksu.edu
Name: Lloyd Willard, EM Lab

Organization: Kansas State University

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Dear All,

Could anyone give me the name of a company that sells/distributes cold labeled transferrine (10nm preferably).

Thank you very much!

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 01:02:38 2005



From: :      Colin.Veitch-at-csiro.au
Date: Fri, 17 Jun 2005 16:01:54 +1000
Subject: [Microscopy] Film thickness measurements with EDXS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,

A colleague has some thin metal films (5-50nm) on Si wafers and we need
to know the relative film thicknesses. We don't need to know the
absolute thickness. Is there any way to use the EDXS signal to get
relative thickness? I.e. use the suppression of the Si signal? Or will
the results be too unreliable? We have no other way here to do the job.

Any suggestions (including don't bother trying it!!!) will be welcome.

Cheers.

Colin Veitch

Electron Microscopist

CSIRO Textile and Fibre Technology

PO Box 21, BELMONT, Vic. 3216. Australia.

E-mail: colin.veitch-at-csiro.au

Web: http://www.tft.csiro.au

Tel: +61 (0) 3 5246 4000
Mob: 0438 538 475
Fax: +61 (0) 3 5246 4811



The information contained in this e-mail message may be privileged or
confidential information. If you are not an intended recipient, you may
not copy, distribute or take any action in reliance on it. If you have
received this message in error, please telephone CSIRO Textile and Fibre
Technology on +61 3 5246 4000.




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 07:55:19 2005



From: Kestutis Smalinskas :      smalinskas-at-yahoo.com
Date: Fri, 17 Jun 2005 05:54:37 -0700 (PDT)
Subject: [Microscopy] Re: Film thickness measurements with EDXS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Colin,

This exercise should be rather straightforward. I've
done it a number of times. If the film is thin
enough, you'll pick up the Si under the metal film.
The thinner the film, the stronger the Si peak. This
will vary with accelerating voltage. Find a voltage
that picks up both the metal film and Si with
reasonably large peaks. To compare the two samples
you'll need to collect a spectrum from each sample
using the same accelerating voltage constant. Then
compare the two spectra looking at the ratio of
metal/Si peak heights within each spectrum.

To take it a step futher, you could get absolute
values if you had control samples with known metal
thickness for comparison.

Stu Smalinskas, P.E.
SKF North American Technical Center
Plymouth, MI
(734) 414-6862
stu.smalinskas-at-skf.com

--- Colin.Veitch-at-csiro.au wrote:

}
}
}
------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The
} Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help
}
http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
-------------------------------------------------------------------------------
}
} Hi,
}
} A colleague has some thin metal films (5-50nm) on Si
} wafers and we need
} to know the relative film thicknesses. We don't
} need to know the
} absolute thickness. Is there any way to use the
} EDXS signal to get
} relative thickness? I.e. use the suppression of the
} Si signal? Or will
} the results be too unreliable? We have no other way
} here to do the job.
}
} Any suggestions (including don't bother trying
} it!!!) will be welcome.
}
} Cheers.
}
} Colin Veitch
}
} Electron Microscopist
}
} CSIRO Textile and Fibre Technology
}
} PO Box 21, BELMONT, Vic. 3216. Australia.
}
} E-mail: colin.veitch-at-csiro.au
}
} Web: http://www.tft.csiro.au
}
} Tel: +61 (0) 3 5246 4000
} Mob: 0438 538 475
} Fax: +61 (0) 3 5246 4811
}
}
}
} The information contained in this e-mail message may
} be privileged or
} confidential information. If you are not an intended
} recipient, you may
} not copy, distribute or take any action in reliance
} on it. If you have
} received this message in error, please telephone
} CSIRO Textile and Fibre
} Technology on +61 3 5246 4000.
}
}
}
}




__________________________________
Yahoo! Mail
Stay connected, organized, and protected. Take the tour:
http://tour.mail.yahoo.com/mailtour.html



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 08:08:03 2005



From: Chaoying Ni :      cni-at-UDel.Edu
Date: Fri, 17 Jun 2005 09:07:15 -0400 (EDT)
Subject: [Microscopy] Re: Film thickness measurements with EDXS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

There are plenty of ways to do this. Immediately coming up in my mind is
the 4-point resistivity measurement with a routine device in any chip
manufacturing shop. x-section SEM should you have a hi-res SEM and be
able to afford for the destruction of the sample or x-section TEM is the
other

****************************************
Chaoying Ni, PhD
The W.M. Keck Electron Microscope Facility
Facility location: 022 Spencer Laboratory
Mailing address:
201 duPont Hall
Dept of Matls Sci & Eng
University of Delaware
Newark, DE 19716
(302) 831-2318(L); -4545(Fax)
http://eml.masc.udel.edu
*****************************************

On Fri, 17 Jun 2005 Colin.Veitch-at-csiro.au wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Hi,
}
} A colleague has some thin metal films (5-50nm) on Si wafers and we need
} to know the relative film thicknesses. We don't need to know the
} absolute thickness. Is there any way to use the EDXS signal to get
} relative thickness? I.e. use the suppression of the Si signal? Or will
} the results be too unreliable? We have no other way here to do the job.
}
} Any suggestions (including don't bother trying it!!!) will be welcome.
}
} Cheers.
}
} Colin Veitch
}
} Electron Microscopist
}
} CSIRO Textile and Fibre Technology
}
} PO Box 21, BELMONT, Vic. 3216. Australia.
}
} E-mail: colin.veitch-at-csiro.au
}
} Web: http://www.tft.csiro.au
}
} Tel: +61 (0) 3 5246 4000
} Mob: 0438 538 475
} Fax: +61 (0) 3 5246 4811
}
}
}
} The information contained in this e-mail message may be privileged or
} confidential information. If you are not an intended recipient, you may
} not copy, distribute or take any action in reliance on it. If you have
} received this message in error, please telephone CSIRO Textile and Fibre
} Technology on +61 3 5246 4000.
}
}
}
}


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 10:19:21 2005



From: Jeff Stewart :      jeff-at-metallography.com
Date: Fri, 17 Jun 2005 11:12:59 -0400
Subject: [Microscopy] International Metallographic Contest

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

It's not too late to enter the International Metallographic Contest and
Exhibit co-sponsored since 1972 by the International Metallographic Society
and ASM International. The contest will be held in conjunction with the 38th
annual technical meeting of the IMS and the M&M 2005 meeting this August in
Honolulu. Twelve categories of competition. Best in show receives $3000 and
the prestigious Jaquet-Lucas Award. Cash awards for first, second, and third
place winners in eleven of the categories. Entries are prominently displayed
during the M&M 2005 meeting and again in the fall during the ASM Annual
Event. Deadline for entries is July 18. For additional information including
rules, tips for creating a winning entry, judging guidelines, and examples
of winning entries contact me or visit
http://www.internationalmetallographicsociety.org/contest.html.

Jeff Stewart
International Metallographic Contest Chair
Metallographic Laboratory Manager
Stern-Leach Co.
49 Pearl Street
Attleboro, MA 02703 USA
Phone: 508-222-7400 extension 1329
FAX: 508-699-4030




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 13:00:34 2005



From: diller-at-stefan-diller.com (by way of MicroscopyListserver)
Date: Fri, 17 Jun 2005 12:59:53 -0500
Subject: [Microscopy] viaWWW: image database of herbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (diller-at-stefan-diller.com) from http://microscopy.com/MLFormMail.html on Thursday, June 16, 2005 at 08:35:16
---------------------------------------------------------------------------

Email: diller-at-stefan-diller.com
Name: Stefan Diller

Organization: Photography

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Dear All,
does anybody know of an image database in the net, preferably freely accessible, which deals with herbs and especially with the microstructure of herbs using SEM and TEM?

Best regards,
Stefan


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 13:02:21 2005



From: Lou Ross :      RossLM-at-missouri.edu
Date: Fri, 17 Jun 2005 13:01:36 -0500
Subject: [Microscopy] Re: Film thickness measurements with EDXS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Colin,

For years I have used a free shareware program called GMRfilm to
measure submicrometer film thicknesses on substrates. It is an old
DOS based program (I think there might be newer programs out now) and
can be found on the MSA website through the reference and educational
link or at http://www.amc.anl.gov. From the parent directory, go to
02-MMSLib/ XEDS/ GMRFILM/.

I routinely use it to calibrate coating thicknesses for our sputter
coater and vacuum evaporator. As Stu mentioned, you need to choose
the appropriate keV to excite both the x-ray signal from the film and
the substrate. Then under identical conditions measure collect
spectra from pure standards. Extract the net peak intensities and
calculate the K-ratios. Once you have the K-ratios and other
operating and specimen parameters are entered into GMRfilm to
calculate the film thickness.

If you want, feel free to contact me off line for more information.

Lou Ross




}
} Hi,
}
} A colleague has some thin metal films (5-50nm) on Si wafers and we need
} to know the relative film thicknesses. We don't need to know the
} absolute thickness. Is there any way to use the EDXS signal to get
} relative thickness? I.e. use the suppression of the Si signal? Or will
} the results be too unreliable? We have no other way here to do the job.
}
} Any suggestions (including don't bother trying it!!!) will be welcome.
}
} Cheers.
}
} Colin Veitch
}
} Electron Microscopist
}
} CSIRO Textile and Fibre Technology
}
} PO Box 21, BELMONT, Vic. 3216. Australia.
}
} E-mail: colin.veitch-at-csiro.au
}
} Web: http://www.tft.csiro.au
}
} Tel: +61 (0) 3 5246 4000
} Mob: 0438 538 475
} Fax: +61 (0) 3 5246 4811
}
}
}
} The information contained in this e-mail message may be privileged or
} confidential information. If you are not an intended recipient, you may
} not copy, distribute or take any action in reliance on it. If you have
} received this message in error, please telephone CSIRO Textile and Fibre
} Technology on +61 3 5246 4000.


--
Senior Electron Microscope Specialist
Electron Microscopy Core Facility
W136 Veterinary Medicine
University of Missouri
Columbia, MO 65211-5120
(573) 882-4777, fax 884=2227
email: rosslm-at-missouri.edu
http://www.emc.missouri.edu


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 15:33:55 2005



From: sschmech-at-u.washington.edu (by way of MicroscopyListserver)
Date: Fri, 17 Jun 2005 15:33:12 -0500
Subject: [Microscopy] viaWWW: strange cytoplasmic changes : reovirus effects on

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (sschmech-at-u.washington.edu) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday, June 17, 2005 at 15:22:09
---------------------------------------------------------------------------

Email: sschmech-at-u.washington.edu
Name: Steve Schmechel

Organization: University of Washington

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: In a study of reovirus effects on infected cells in culture, we are finding strange cytoplasmic changes. Cells infected with some strains (and not others) have strange cytoplasmic structures. These are non-membrane bound clear spaces that appear as wide splits in the cytoplasm. They are large, approximatly the length of the nucleus with a width approximately one fifth of length. They "smile" at the middle and are curved. Although none are directly next to the nucleus, they take a curved form that generally follows a similar curve as the nuclear membrane. In most cells there are many of these. The result is that the cells have a zebra stripe appearance due to these multiple splits in the cytoplasm.

Has anyone ever seen something like this. I'm happy to email you a sample if you like.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 17 16:33:23 2005



From: Edward Calomeni :      calomeni-1-at-medctr.osu.edu
Date: Fri, 17 Jun 2005 17:32:13 -0400
Subject: [Microscopy] Re:reovirus effectson infected cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Steve,

Without an image to study, my first guess is that the 'inclusions' are
cholesterol clefts.
Frank Ramig, at Baylor College of Medicine, Division of Biochemical
Virology has many, many years under his belt in working with reoviruses.
He might be a good person to ask about these things.

Best of luck,

Ed

Edward Calomeni
Director EM Lab
Ohio State University - Pathology
M018 Starling Loving Hall
320 W. 10th Ave.
Columbus, OH 43210-1240
614-293-5580 (office)
614-293-8806 (lab)
calomeni-1-at-medctr.osu.edu

} } } by way of MicroscopyListserver {sschmech-at-u.washington.edu}
6/17/2005 4:33 PM } } }


------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of
America

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (sschmech-at-u.washington.edu) from
http://microscopy.com/MicroscopyListserver/MLFormMail.html on Friday,
June 17, 2005 at 15:22:09
---------------------------------------------------------------------------

Email: sschmech-at-u.washington.edu
Name: Steve Schmechel

Organization: University of Washington

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: In a study of reovirus effects on infected cells in culture,
we are finding strange cytoplasmic changes. Cells infected with some
strains (and not others) have strange cytoplasmic structures. These are
non-membrane bound clear spaces that appear as wide splits in the
cytoplasm. They are large, approximatly the length of the nucleus with
a width approximately one fifth of length. They "smile" at the middle
and are curved. Although none are directly next to the nucleus, they
take a curved form that generally follows a similar curve as the nuclear
membrane. In most cells there are many of these. The result is that
the cells have a zebra stripe appearance due to these multiple splits in
the cytoplasm.

Has anyone ever seen something like this. I'm happy to email you a
sample if you like.

---------------------------------------------------------------------------





From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 08:17:28 2005



From: Gervais.Sawyer-at-bcuc.ac.uk (by way of Ask-A-Microscopist)
Date: Mon, 20 Jun 2005 08:16:47 -0500
Subject: [Microscopy] AskAMicroscopist: SEM image OF ash structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (Gervais.Sawyer-at-bcuc.ac.uk) from http://www.microscopy.org/Ask-A-Microscopist/Ask-A-Microscopist.html on Monday, June 20, 2005 at 04:56:39
---------------------------------------------------------------------------

Email: Gervais.Sawyer-at-bcuc.ac.uk
Name: Gervais Sawyer

Organization: BCUC, UK

Education: Graduate College

Location: High Wycombe, Buckinghamshire, UK

Question: I am trying to SEM image an ash structure after burning pieces of wood. These are extremely beautiful and could reveal some microstructure. However, the ash is very delicate and disintegrates when I go to vacuum. The same thing happens when I try to sputter coat. Any ideas please?

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 08:53:09 2005



From: Kerrie Venner :      k.venner-at-ion.ucl.ac.uk
Date: Mon, 20 Jun 2005 14:52:19 +0100
Subject: [Microscopy] TEM digitisation: digital plates vs digital camera systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have been given the unenviable task of having to decide between a digital
camera system and a digital imaging plate system (the Ditabis system) to
enable our older TEMs to become fully digital imaging systems. We have Jeol
1200EX and 100CX instruments and a Phillips CM10. Has anyone had experience
of the Ditabis system? In the UK there is only one currently in use, and I
have contact with that person. Our samples are biological samples, some
routine clinical specimens, but also research samples and cell culture
samples. We currently use Kodak em film 4489 and require the replacement
system to have comparable resolution.
Any comments would be gratefully received.






From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 11:40:01 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 20 Jun 2005 09:38:27 -0700
Subject: [Microscopy] Re: TEM digitisation: digital plates vs digital camera systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Jun 20, 2005, at 6:52 AM, Kerrie Venner wrote:

} We currently use Kodak em film 4489 and require the replacement
} system to have comparable resolution.
} Any comments would be gratefully received.

Dear Kerrie,
The grain size on 4489 is extremely small (~1 um), so the resolution
is much better than can be achieved with any image plate system. Image
plates have many advantages over film, such as linearity and ease of
digitization, so they are better than film for quantitation, but for
enlargement by many times or other processes that require high
resolution, film is still superior. I do not know any specifics about
the Ditabis system.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 11:58:31 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Mon, 20 Jun 2005 09:57:35 -0700
Subject: [Microscopy] TEM digitisation: digital plates vs digital camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Kerrie;
There is an operational difficulty with imaging plate systems in
that magnification and serial number are not recorded on the plate. In
fact, nothing is recorded on the plate except the electron image. Think
about getting one plate out of order, and how that might foul up your
operations.

John Mardinly
Intel

This opinion is that of the author and does not represent the opinion of
Intel corporation.

-----Original Message-----
} From: Kerrie Venner [mailto:k.venner-at-ion.ucl.ac.uk]
Sent: Monday, June 20, 2005 6:52 AM
To: Microscopy-at-microscopy.com

I have been given the unenviable task of having to decide between a
digital
camera system and a digital imaging plate system (the Ditabis system) to
enable our older TEMs to become fully digital imaging systems. We have
Jeol
1200EX and 100CX instruments and a Phillips CM10. Has anyone had
experience
of the Ditabis system? In the UK there is only one currently in use,
and I
have contact with that person. Our samples are biological samples, some
routine clinical specimens, but also research samples and cell culture
samples. We currently use Kodak em film 4489 and require the
replacement
system to have comparable resolution.
Any comments would be gratefully received.








From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 13:05:13 2005



From: Hong Yi :      hyi-at-emory.edu
Date: Mon, 20 Jun 2005 14:04:31 -0400
Subject: [Microscopy] JB-4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Can anyone tell me if JB-4 embedded sample can be sectioned on an
ultramicrotome (Leica Ultracut S), and what the appropriate thickness
should it be cut at? Thank you in advance.

HONg
Emory



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 13:24:48 2005



From: Tobias Baskin :      baskin-at-bio.umass.edu
Date: Mon, 20 Jun 2005 14:24:03 -0400
Subject: [Microscopy] Re: JB-4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hong,
I think JB-4 is a species of methacrylate resin so sure you
can cut it on an Ultra. I'd guess anywhere from 2 um semis right down
to 60nm ultras (though it may be difficult to get all the way that
thin) depending on what you want to do.

Happy slicing,
Tobias

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
_ ____ __ ____
/ \ / / \ / \ \ Tobias I. Baskin
/ / / / \ \ \ Biology Department
/_ / __ /__ \ \ \__ 611 N. Pleasant St.
/ / / \ \ \ University of Massachusetts
/ / / \ \ \ Amherst, MA, 01003
/ / ___ / \ \__/ \ ____
http://www.bio.umass.edu/biology/baskin/
Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 13:40:10 2005



From: Tom Phillips :      phillipst-at-missouri.edu
Date: Mon, 20 Jun 2005 13:39:23 -0500
Subject: [Microscopy] Re: JB-4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have cut 0.5 to 2 um thick JB-4 sections using a dry glass knife without
a boat on many occasions. As the section starts to come off on the knife
face, help it along with one of the tines of a fine forceps so that the
section doesn't crumble up. Gently lift the cut section and drop in one
motion on a drop of water. be sure to let go of the section before it
touches the water or it will shrivel up against the forceps. Heat the
section down on a hot plate as one would do for an Epon section. It works
but is tedious. Furthermore, I would say the frustration is unnecessary
since BMMA (butyl-methyl methacrylate/methyl methacrylate mixtures) are
less expensive (generic ingredients), easier to cut (cuts on water filled
boats), and can be deplasticized using acetone. We love BMMA and haven't
used JB-4 ever since the esteemed Dr. Tobias Baskin told us about it. I
don't have the reference to his paper but he monitors this listserver and
perhaps will post it.



At 01:04 PM 06/20/05, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thomas E. Phillips, PhD
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400

573-882-4712 (office)
573-882-0123 (fax)
PhillipsT-at-missouri.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 13:45:12 2005



From: Brendan J Griffin :      bjg-at-cmm.uwa.edu.au
Date: Mon, 20 Jun 2005 13:44:28 -0500
Subject: [Microscopy] Positon Open : nanoSIMS 50 ion probe position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html




Dear all

Please draw the following position description to the attention of
any potentially interested people. We can be flexible for the right
person, within usual constraints. We have just won a new WA State
Government Centre of Excellence grant that will provide an additional
position into the nanoSIMS laboratory later in the year or early next
year.

Advertised in: The West Australian (Prof Appts) Saturday 11 June 2005
The Australian (HES) Wednesday 15 June 2005


RESEARCH OFFICER/SENIOR RESEARCH OFFICER (REF: 885)
CENTRE FOR MICROSCOPY AND MICROANALYSIS

o Ongoing appointment
o Salary range: HEE Level 6 $50,795 - $53,762 p.a.
o Salary range: HEE Level 7 $55,739 - $62,000 p.a.
o Closing date: Tuesday, 28 June 2005

A CAMECA nanoSIMS 50 high-resolution ion microprobe was installed in
the Centre for Microscopy and Microanalysis (CMM) at The University
of Western Australia, as part of the NANO-MNRF, in June, 2003. The
NANO-MNRF (www.nano.org.au) is a Major National Research Facility
that links advanced nano-scale characterisation equipment at the
Universities of Sydney, New South Wales, Queensland, Western
Australia (UWA) and Melbourne. The range of CMM microscopy
instrumentation and expertise is extensive (see
http://cmm.uwa.edu.au) such that the centre and MNRF represent a
premium environment for research services, research training and
research programs. UWA is also a major partner in local research
consortia in isotope science with a range of stable and radiogenic
isotope facilities, including two SHRIMP-II.

We are seeking a highly motivated, self-guided scientist who has the
ability to work with other staff within the Centre and its users. The
prime responsibility will be to manage a CAMECA nanoSIMS 50 ion
microprobe as a world-class National Facility. The position will have
the core role in a strong team led locally by Associate Professor
Brendan Griffin and nationally by Professor Simon Ringer (NANO MNRF
Executive Director). This position is advertised as a Level 6 or 7
depending on qualifications and experience. The minimum qualification
is a relevant degree or equivalent experience.

Application Details: Interested applicants must obtain the
application package and address the prerequisites and selection
criteria. These essential details can be accessed from the vacancy
page on http://jobs.uwa.edu.au/ or the 24 hour "hotline" on 6488
3733. To discuss or clarify any aspects of the position please
contact the director of the CMM, Associate Professor Brendan Griffin
on (08) 6488 2770 or email bjg-at-cmm.uwa.edu.au.

Committed to recruiting, developing and retaining the highest quality staff

--




Brendan J. Griffin
Director & Assoc. Professor in Electron Microscopy
Centre for Microscopy and Microanalysis (M010)
Director Western Australian Centre for Microscopy
Associate Director NANO-MNRF
President Australian Microbeam Analysis Society
The University of Western Australia
First floor, Physics Building,
35 Stirling Highway
CRAWLEY, WA, AUSTRALIA 6009
ph 61-8-6488-2739 fax 6488-1087
mobile 0409-104-096
bjg-at-cmm.uwa.edu.au
http://cmm.uwa.edu.au/

CRICOS Provider No. 00126G


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 13:57:44 2005



From: diller-at-stefan-diller.com (by way of MicroscopyListserver)
Date: Mon, 20 Jun 2005 13:57:02 -0500
Subject: [Microscopy] viaWWW: Old KEVEX detector on KEVEX DELTA 8000

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (diller-at-stefan-diller.com) from http://microscopy.com/MLFormMail.html on Monday, June 20, 2005 at 09:55:20
---------------------------------------------------------------------------

Email: diller-at-stefan-diller.com
Name: Stefan Diller

Title-Subject: [Microscopy] [Filtered] Old KEVEX detector on KEVEX DELTA 8000

Question: Dear All,
maybe there is someone out there who can help me connect an old KEVEX EDS (seems to be Model 3801M?) detector with a Preamplifier Modell 2002 on my "new" KEVEX Delta 8000 unit.

I especially need the layouts of
1- the PREAMP1 connector on the rear of the KEVEX 8000 concerning voltage output and preamp output coming from the detector (see image at: http://www.stefan-diller.com/rem/Kevex1.jpg)
2- the POWER connector on the EDS detector and also the voltages needed for the preamp Modell 2002 in the detector
(see image at: http://www.stefan-diller.com/rem/Kevex2.jpg and http://www.stefan-diller.com/rem/Kevex3.jpg )
3- or the simple answer: Is it possible to use this detector on the KEVEX 8000?


Best regards,
Stefan Diller

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 14:11:37 2005



From: Tobias Baskin :      baskin-at-bio.umass.edu
Date: Mon, 20 Jun 2005 15:46:17 -0400
Subject: [Microscopy] Re: JB-4 (BMM recips)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello All:

Sort of a way out idea but what about vapor impregnation by Cyanoacrylates
then sputter coat? Like Forensic scientists use to get fingerprints of
unusual places.

Cheers!

Al Coritz
Electron Microscopy Sciences
www.emsdiasum.com


----- Original Message -----
} From: "by way of Ask-A-Microscopist" {Gervais.Sawyer-at-bcuc.ac.uk}
To: {microscopy-at-microscopy.com}
Sent: Monday, June 20, 2005 9:16 AM

Group,
As Tom mentioned, mixtures of butyl and methyl and
methacrylate produce easy to section blocks that can serve a variety
of purposes. These mixtures predate epoxies for TEM use and they are
unstable in TEM beams. But the other side of that coin is most of the
resin can be extracted after sectioning with a 10 minute incubation
in aceteone. This provides excellent access for an antibody probe to
most of the section volume. So for light level immuno work, BMM is
great stuff. My innovation was to dissolve some DTT in the resin mix
which keeps those pesky free radicals from oxidizing the daylights
out of the sample.

My original paper on this is:
Baskin TI, Busby CH, Fowke LC. Sammut M, Gubler F (1992)
Improvements in immunostaining samples embedded in methacrylate:
Localization of microtubules and other antigens throughout developing
organs in plants of diverse taxa. Planta 187: 405 - 413.

And a few later findings in the methods section of this paper:

Baskin TI, Wilson JE (1997) Inhibitors of protein kinases and
phosphatases alter root morphology and disorganize cortical
microtubules. Plant Physiology 113: 493 - 502.

If anyone wants to try this, contact me off line and I can
provide a few extra resources for you.

Happy slicing,
Tobias


}
} I have cut 0.5 to 2 um thick JB-4 sections using a dry glass knife
} without a boat on many occasions. As the section starts to come off
} on the knife face, help it along with one of the tines of a fine
} forceps so that the section doesn't crumble up. Gently lift the cut
} section and drop in one motion on a drop of water. be sure to let go
} of the section before it touches the water or it will shrivel up
} against the forceps. Heat the section down on a hot plate as one
} would do for an Epon section. It works but is tedious. Furthermore,
} I would say the frustration is unnecessary since BMMA (butyl-methyl
} methacrylate/methyl methacrylate mixtures) are less expensive
} (generic ingredients), easier to cut (cuts on water filled boats),
} and can be deplasticized using acetone. We love BMMA and haven't
} used JB-4 ever since the esteemed Dr. Tobias Baskin told us about
} it. I don't have the reference to his paper but he monitors this
} listserver and perhaps will post it.
}
}
}
} At 01:04 PM 06/20/05, you wrote:
}
} } ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America


--
_ ____ __ ____
/ \ / / \ / \ \ Tobias I. Baskin
/ / / / \ \ \ Biology Department
/_ / __ /__ \ \ \__ 611 N. Pleasant St.
/ / / \ \ \ University of Massachusetts
/ / / \ \ \ Amherst, MA, 01003
/ / ___ / \ \__/ \ ____
http://www.bio.umass.edu/biology/baskin/
Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 20:03:50 2005



From: Stacey.Andringa-at-uc.edu (by way of MicroscopyListserver)
Date: Mon, 20 Jun 2005 20:03:07 -0500
Subject: [Microscopy] viaWWW: Siemens IA Filaments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (Stacey.Andringa-at-uc.edu) from http://www.microscopy.com/MLFormMail.html on Monday, June 20, 2005 at 14:35:52
---------------------------------------------------------------------------

Email: Stacey.Andringa-at-uc.edu
Name: Stacey Andringa

Organization: University of Cincinnati

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I need to have filaments retipped for a Siemens IA. Can anyone recommend a vendor for this?
Thanks.

Stacey Andringa

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 20:05:09 2005



From: artcollc-at-emirates.net.ae (by way of Ask-A-Microscopist)
Date: Mon, 20 Jun 2005 20:04:26 -0500
Subject: [Microscopy] AskAMicroscopist: Oversized Petrographic Thin Section Coverslips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (artcollc-at-emirates.net.ae) from on Monday, June 20, 2005 at 09:36:41
---------------------------------------------------------------------------

Email: artcollc-at-emirates.net.ae
Name: k. j. dhar

Organization: artco llc.

Education: Graduate College

Location: Dubai, United Arab Emirates.

Question: We are a marketing company, distributing (since 1989)all types of Material Testing equipment.

Currently, one of our customers (who is engaged in Oil production, refining, distribution) has set up a Petroleum Institute. They have a requirement for Oversized Petrographic Thin Section Coverslips, in two sizes 42mmx25mm and
74x49mm with 0.16mm thickness.

We seek your assistance/suggestions on the likely source from whom this can be ordered.

Thanking you in anticipation and with best regards,

Kiran Dhar


---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 20 21:09:52 2005



From: Barbara :      bfoster-at-mme1.com
Date: Mon, 20 Jun 2005 21:08:51 -0500
Subject: [Microscopy] Re: AskAMicroscopist: SEM image OF ash structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Gervais,

Some years ago, Buehler Instruments (Waukegan,IL), made a special device to mount delicate samples like this. It consisted of a specially constructed vacuum jar in which was mounted a small ladle that could pour epoxy into mounting cups. The specimen was put in the cups, the jar closed, the epoxy poured, then a vacuum drawn. The result: the epoxy penetrated the delicate structure and permitted microtoming.

I actually saw the ash on a cigar mounted this way, so think that it should be a good solution for your problem. You can probably access Buehler on the web. If not, contact me off line and I will see if I can find their contact info.

Good hunting!

Barbara Foster

Microscopy/Microscopy Education
313 S Jupiter Rd, Suite 100
Allen, TX 75002
P: 972-954-8011
W: www.MicroscopyEducation.com

P. S.
Need a good general reference or light microscopy text? Call us today to learn more about "Optimizing LIght Microscopy". Copies still available through MME... even for class-room lots ... and we give quantity discounts. Call us at (972)954-8011.

At 08:16 AM 6/20/2005, Gervais.Sawyer-at-bcuc.ac.uk wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 02:45:26 2005



From: gillian.2.brown-at-gsk.com
Date: Tue, 21 Jun 2005 08:43:56 +0100
Subject: [Microscopy] Re: JB-4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,
it depends on what microscopy you are actually doing, replies already in
are for TEM.
I am using a modified version of JB-4 (much less catalyst and 'B') for IHC
at the light microscope level and the resin does not need to be removed. I
cut 2micron sections on a dry knife (often the same one I trimmed on as
the resin is so soft) but I can go lower. No I can't do TEM off the same
block..
Refs: Casey et al. (1988) Am J Pathol;131:183-189
Britten et al. (1993) Biotech Histochem;68:271-280

Gill Brown
Histopathology Group
Asthma and Allergy Disease Biology
GlaxoSmithKline Medicines Research Centre,
UK




"Hong Yi" {hyi-at-emory.edu}
20-Jun-2005 19:04

To
" {microscopy-at-microscopy.com} {microscopy-at-microscopy.com} "
cc

Subject
JB-4








------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Can anyone tell me if JB-4 embedded sample can be sectioned on an
ultramicrotome (Leica Ultracut S), and what the appropriate thickness
should it be cut at? Thank you in advance.

HONg
Emory








From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 02:50:24 2005



From: Olivia Lowery :      xejkzaluzec-at-microscopy.com
Date: Tue, 21 Jun 2005 03:49:38 -0500
Subject: [Microscopy] Offering Purchases for Americans

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Good day,

Homeowner

You have been pre-approved for a $447,633 Home Loan at a 3.94 Fixed Rate.
This offer is being extended to you unconditionally and your credit is in no way a factor.

To take Advantage of this Limited Time opportunity

All we ask is that you visit our Website and complete
The 1 minute post Approval Form

http://yIwVLx.max-loans.com/1/index/ryn/DFmCmcY

Best Wishes,

Olivia Lowery


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 05:11:26 2005



From: Robert H. Olley :      hinmeigeng-at-hotmail.com
Date: Tue, 21 Jun 2005 10:10:42 +0000
Subject: [Microscopy] Philips SEM 515 available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We have a Philips SEM 515, about 17 years old, available. The instrument
itself is in very good working order, apart from a small electrical problem
with the pumping system which we did not bother to have fixed since we moved
to a newer instrument.

Anyone interested?

-----------------------------------
Robert H. Olley
Reply to: R.H.Olley-at-reading.ac.uk
URL: http://www.rdg.ac.uk/~spsolley
-----------------------------------




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 12:22:27 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Tue, 21 Jun 2005 10:21:38 -0700
Subject: [Microscopy] Final Epitaph for film-Kodak Ends Production of B/W print paper

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



This announcement in Yahoo News:
http://news.yahoo.com/s/ap/20050615/ap_on_bi_ge/kodak_paper


John Mardinly
Intel





From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 17:49:44 2005



From: gardnere-at-wpunj.edu (by way of MicroscopyListserver)
Date: Tue, 21 Jun 2005 17:49:03 -0500
Subject: [Microscopy] viaWWW: Confocal microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (gardnere-at-wpunj.edu) from http://www.microscopy.com/MLFormMail.html on Tuesday, June 21, 2005 at 13:50:12
---------------------------------------------------------------------------

Email: gardnere-at-wpunj.edu
Name: Eileen Gardner

Organization: William Paterson University

Title-Subject: [Microscopy] [Filtered] MListserver: Confocal microscope

Question: I am the chair of a Biology department at a state university in NJ. I am planning on purchasing a confocal microscope which would be used for both teaching and research. Do you have any ideas as to which brands would be easiest to use and maintain? There will be no technician dedicated to this microscope. Thanks.

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 18:37:37 2005



From: Jay Campbell :      microtomy-at-gmail.com
Date: Tue, 21 Jun 2005 18:36:55 -0500
Subject: [Microscopy] TEM digital camera options

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Listers,
We are trying to find a workable solution for a problem that I
believe others may have run up against regarding digital image
acquisition. We have a Tecnai T12 microscope with a high resolution
bottom mount digital camera (Gatan 4K x 4K). The problem is due to
the slow frame rate and small field of view of the camera makes it
difficult to focus and capture low magnification survey shots. We
have considered
adding a side-mount CCD camera to complement the main camera but would
like to hear from other users to see what our options are. Does anyone
have experience with direct video rate cameras (side mounted) for the
purpose of locating and focusing areas of interest?

Another option would be to add a multiple image stitching software
to the current camera and merge several higher mag images. My
concern here is image/ specimen drift over the time these multiple
acquisitions are made. I work primarily with slot grids and support
film which tends to exacerbate such problems.

Thanks in advance,
Jay


Jay Campbell
Research Specialist
University of Wisconsin
Laboratory of Molecular Biology
R.M. Bock Labs
1525 Linden Drive
Madison, WI 53706
jmcampbe-at-wisc.edu
608 263 8481



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 18:43:21 2005



From: Gordon Vrololjak :      gvrdolja-at-nature.berkeley.edu
Date: Tue, 21 Jun 2005 16:42:31 -0700 (PDT)
Subject: [Microscopy] Re: viaWWW: Confocal microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I'd say there should be a technician or your going to have some trouble
with getting it to work. Maybe appoint someone to take care of it and
train people. Go with zeiss, they have nice confocal systems that are
really easy to use and integrate well with the software for
control/imaging.
Gordon

\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\
Gordon Ante Vrdoljak Electron Microscope Lab
AOL/IM rakhasha http://nature.berkeley.edu/~gvrdolja 26 Giannini Hall
gvrdolja-at-nature.berkeley.edu UC Berkeley
phone (510) 642-2085 Berkeley CA 94720-3330
fax (510) 643-6207 cell (510) 290-6793

On Tue, 21 Jun 2005, by way of MicroscopyListserver wrote:

}
}
} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
} Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (gardnere-at-wpunj.edu) from http://www.microscopy.com/MLFormMail.html on Tuesday, June 21, 2005 at 13:50:12
} ---------------------------------------------------------------------------
}
} Email: gardnere-at-wpunj.edu
} Name: Eileen Gardner
}
} Organization: William Paterson University
}
} Title-Subject: [Microscopy] [Filtered] MListserver: Confocal microscope
}
} Question: I am the chair of a Biology department at a state university
} in NJ. I am planning on purchasing a confocal microscope which would be
} used for both teaching and research. Do you have any ideas as to which
} brands would be easiest to use and maintain? There will be no
} technician dedicated to this microscope. Thanks.
}
} ---------------------------------------------------------------------------
}


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 20:36:24 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Tue, 21 Jun 2005 18:36:18 -0700
Subject: [Microscopy] Re: TEM digital camera options

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Jay
I would think, the good solution to you would be side-mounted camera with
optical coupling - they normally shows quite decent refresh rate. Perhaps,
it would be wise to use the same manufacturer for 100% compatibility with
existing camera. As for focusing problem - it's kind of strange. Gatan's
4x4 has view area compatible to the film. As for "slowness", yes, it's
slow. You need to use at least 4x (even 8x) binning in the "view" mode. I
do find that focusing with wobbler is extremely effective with my camera -
BioScan 600W - it's top-mounted, so comparison is not quite adequate. They
have "loupe" in DM, which is also helpful. Gatan also offered auto-focus
plug-in and as far as I remember it was working nicely at their
demonstration with 4x4 camera. DigitalMontage is also great - they used
beam shift to extend the view, so you could take pictures quite quickly,
like 3x3 or even 4x4 arrays. If you'll evaporate carbon on top of your
plastic film - it significantly reduces the drift. You may do it over the
sections (even better). Sergey

At 04:36 PM 6/21/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From MicroscopyL-request-at-ns.microscopy.com Tue Jun 21 20:48:31 2005



From: Joe Kulik :      juk12-at-psu.edu
Date: Tue, 21 Jun 2005 22:09:47 -0400
Subject: [Microscopy] TEM: Cone angle of LaB6 emitter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Here are some thoughts from my own experience.

} From what I have heard, the optics of the major instruments are all
comparable. The choice of instrument, then, depends on ease of
operation (software), reliability, and service. There are some
technical details in the design of the spectral features that might
bear on the sensitivity of the instrument, though.

I would recommend that you think carefully about what you want to do
with this instrument. Do you want an upright or inverted microscope?
You should decide which wavelengths you need for the stains you are
likely to encounter, and make a similar analysis of the lenses that
you will need (long or short working distances,magnifications, etc)..
You should also decide if you need spectral capabilities on both the
excitation and emission detectors.

A confocal microscope is a significant instrument, with a complexity
not that different from the old electron microscopes. You can't just
walk in and use it without training. As Gordon suggested, you shoud
make sure that there is someone with responsibility for the
instrument. This person should be familiar enough to give
instruction and help to individual users.


The software is critical. It isn't that the different software sets
will give you different capabilities, but the ease of different
operations may vary. As a result, you will find that you may have a
different personal reaction to the interface. You, or the person who
will be in charge of the instrument, should sit down with each of the
models that you are considering, and work with the instrument for an
extended period --just looking at the instruments at a meeting, for
instance, will not work. You should also see whether the output
files can be read by 3d party software, such as Metamorph, or ImageJ,
or even Photoshop so that you don't tie up a microscope by doing
image processing with it.

In order to make a decision about the brand, it seems to me that,
given your support situation, you should investigate the service that
is available, and its cost. Service contracts can be quite
expensive, and I have heard anecdotes about the difficulty of service
with most of the manufacturers (although I have also heard some
praises for the same service organizations).

For more details on this, scan the archives of the confocal listserv:
CONFOCAL-at-LISTSERV.BUFFALO.EDU


Hope these musings are helpful. I can give you specific experiences
offlist.

Joel


Date sent: Tue, 21 Jun 2005 17:49:03 -0500
To: microscopy-at-microscopy.com
} From: gardnere-at-wpunj.edu (by way of MicroscopyListserver)

On our JEM-2010 we use a LaB6 emitter with a 90 degree cone and 15
micron tip.

We are considering using a smaller cone angle and/or smaller tip radius.

Does anyone have any comments concerning whether the supposed
improvement in brightness and coherence is worth the shortened lifetime?

_________________________________

Joseph Kulik
Research Associate
Materials Research Institute
The Pennsylvania State University
194 MRI Bldg
University Park, PA 16802

Telephone: 814-865-0344
Fax: 814-863-8561
e-mail: juk12-at-psu.edu
_________________________________




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 08:05:28 2005



From: xiaohutang-at-gmail.com (by way of MicroscopyListserver)
Date: Wed, 22 Jun 2005 08:04:46 -0500
Subject: [Microscopy] viaWWW: 3D image reconstruction software for stereo SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (xiaohutang-at-gmail.com) from http://www.msa.microscopy.org/MicroscopyListserver/MLFormMail.html on Wednesday, June 22, 2005 at 02:27:04
---------------------------------------------------------------------------

Email: xiaohutang-at-gmail.com
Name: Xiaohu Tang

Organization: TU Delft, the Netherlands

Title-Subject: [Microscopy] [Filtered] 3D image reconstruction software for stereo SEM

Question: Hello everyone,

I am looking for a 3D image reconstruction software for stereo SEM to measure the sample volume. It looks like the MeX from Alicona is the only choice so far but the price is too expensive. Could anybody give me an idea for a cheaper software or a better method to measure the volume by SEM? Thank you.

Xiaohu Tang
Microlab
CiTG of Delft Technology University
The Netherlands
x.tang-at-citg.tudelft.nl

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 09:11:18 2005



From: Phaedra McGuinness :      scanning-at-fams.org
Date: Wed, 22 Jun 2005 10:10:36 -0400
Subject: [Microscopy] SCANNING 2006 Tuesday April 25 through Thursday April 27 in Washington, D.C.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Listers,

Just a quick calendar item and announcement regarding SCANNING 2006.
The 17th annual meeting on the scanning microscopies will take place
Tuesday, April 25 through Thursday, April 27 in Washington, D.C. The
Welcome Reception will take place on Tuesday, April 25 at the National
Press Club.

SPECIAL EVENT:
On Wednesday, April 26 there will be a program day at NIST.
Transportation will be provided. This is a joint program day included
in your SCANNING registration fee.

For more information:

SCANNING/FAMS
P.O. Box 485
Mahwah, NJ 07430-0485
Tel: (201) 818-1010
Fax: (201) 818-0086
e-mail: scanning-at-fams.org
Internet: www.scanning.org

See you at the meeting!


Best regards,


Phaedra McGuinness
Managing Editor
SCANNING, The Journal of Scanning Microscopies
P.O. Box 485
Mahwah, NJ 07430
Tel: (201) 818-1010 * Fax: (201) 818-0086 *email: scanning-at-fams.org *
Web: www.scanning.org




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 10:26:52 2005



From: Gordon Couger :      gcc-at-couger.com
Date: Wed, 22 Jun 2005 10:26:01 -0500
Subject: [Microscopy] Re: viaWWW: 3D image reconstruction software for stereo

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi Xiaohu
Have you looked at ImageJ with the needed plugins?

ImageJ's main page is http://rsb.info.nih.gov/ij/

3D Toolkit Plug ins
http://ij-plugins.sourceforge.net/ij-3D-toolkit.html

3D Toolkit is a set of plug ins for 3D and 2D operations on
images in Image/J. The first group of plug ins reads and writes
images in VTKand MetaImage format.
Metaimage Reader
MetaImage Writer
VTK Reader
VTK Writer
The second group of plugins are early prototypes of plugins for:
Morphological dilation (max)
Morphological erosion (min)
Connected threshold region growing
Auto volume clipping

This may be the same set of plug ins
http://ij-plugins.sourceforge.net/3d-toolkit/

A Google search for:
imagej 3d volume
brings over 600 hits and ImageJ and most plugins are free.

ImageJ is Java port with a great deal of enhancement of Sicon
Image that USA National Institute of Health had done and put in
the public domain. Primarily as health sciences imaging tool but
with the open source pugin, scrip an macro language many
disciplines are using it for many things. It is platform
independent running on any machine ruining Java, Unix, Solaris,
Windows, Mac, Mac OS-10, Linux and some more. For a P code
machine it amazingly fast. I would like to see it run on
computer that ran Java in native mode.

It is bit harder to use that Paint shop but new functionality is
available almost on a daily basis and the support from the folks
that write the plug ins is on the whole much much better than
any commercial product that doesn't cost a fortune for
maintenance to provide that kind of service. In many ways ImageJ
support is better than the best commercial service because you
work with the person that wrote the code. It is not uncommon for
someone on the plugin mailing list to help you modify a plugin
for you needs. There are no commercial vendors that can turn out
code this way.

I have worked for software companies as a programmer and letting
the customer work with the programmer is just not done. It uses
too much of the programmers time and the customer finds out more
than management wants them to know about the product. Because
the programmers don't pretend that the patched up out of date
software that most products are made out of are anything else
but what it is. Most of the time management doesn't have the
money to write new software from scratch or look at Cisco when
they tried to write a new operating system for their routers and
got in disagreement with the programmers and the programmers
quit and went into business as Juniper who is now Ciscoes'
biggest competitor in their high end most profitable business.

Source code is available for most plugins and ImageJ will
compile them giving you a complete tool for learning to write
plugins if you want to. Most of use will be content to play the
sorcerer's apprentice and charge a few lines to make a plugin
that is almost what we do just what we want and not become full
blown programmers. But that is very valuable option when you
need something slightly different than the person that developed
the plug in needed. A Google search for 'imagej plugin' brings
up almost 10,000 hits I am sure many of those are duplicates but
3,000 to 5,000 plugins is not an unreasonable estimate.

Best Regards
Gordon
Gordon Couger

I collect links on information related to light microscopes.
www.couger.com/microscope/links/gclinks.html
Please forward anything you think might be useful to others.
Microscope Documentation is at www.science-info.org

by way of MicroscopyListserver wrote:
}
} Email: xiaohutang-at-gmail.com
} Name: Xiaohu Tang
}
} Organization: TU Delft, the Netherlands
}
} Title-Subject: [Microscopy] [Filtered] 3D image
reconstruction software for stereo SEM
}
} Question: Hello everyone,
}
} I am looking for a 3D image reconstruction software for
stereo SEM to measure the sample volume. It looks like the MeX
from Alicona is the only choice so far but the price is too
expensive. Could anybody give me an idea for a cheaper software
or a better method to measure the volume by SEM? Thank you.
}
} Xiaohu Tang
} Microlab
} CiTG of Delft Technology University
} The Netherlands
} x.tang-at-citg.tudelft.nl
}



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 11:50:45 2005



From: Xiaohu Tang :      xiaohutang-at-gmail.com
Date: Wed, 22 Jun 2005 18:50:01 +0200
Subject: [Microscopy] Re: viaWWW: 3D image reconstruction software for stereo SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Gordon, thank you for the valuable information. I checked ImageJ and
could not find the proper plugin for my problem: I want to use the SEM
stereo image pairs by tilting the sample and find out the volume data
from the 3D reconstruction. Based on my understanding (I am not so
familar with this area), most of the ImageJ plugins about 3D volume
depend on a series of image slices but not like my case. All the
suggestions and information will be highly appreciated.

Xiaohu Tang
Microlab
CiTG of Delft Technology University
The Netherlands
x.tang-at-citg.tudelft.nl

On 6/22/05, Gordon Couger {gcc-at-couger.com} wrote:
} Hi Xiaohu
} Have you looked at ImageJ with the needed plugins?
}
} ImageJ's main page is http://rsb.info.nih.gov/ij/
}
} 3D Toolkit Plug ins
} http://ij-plugins.sourceforge.net/ij-3D-toolkit.html
}
} 3D Toolkit is a set of plug ins for 3D and 2D operations on
} images in Image/J. The first group of plug ins reads and writes
} images in VTKand MetaImage format.
} Metaimage Reader
} MetaImage Writer
} VTK Reader
} VTK Writer
} The second group of plugins are early prototypes of plugins for:
} Morphological dilation (max)
} Morphological erosion (min)
} Connected threshold region growing
} Auto volume clipping
}
} This may be the same set of plug ins
} http://ij-plugins.sourceforge.net/3d-toolkit/
}
} A Google search for:
} imagej 3d volume
} brings over 600 hits and ImageJ and most plugins are free.
}
} ImageJ is Java port with a great deal of enhancement of Sicon
} Image that USA National Institute of Health had done and put in
} the public domain. Primarily as health sciences imaging tool but
} with the open source pugin, scrip an macro language many
} disciplines are using it for many things. It is platform
} independent running on any machine ruining Java, Unix, Solaris,
} Windows, Mac, Mac OS-10, Linux and some more. For a P code
} machine it amazingly fast. I would like to see it run on
} computer that ran Java in native mode.
}
} It is bit harder to use that Paint shop but new functionality is
} available almost on a daily basis and the support from the folks
} that write the plug ins is on the whole much much better than
} any commercial product that doesn't cost a fortune for
} maintenance to provide that kind of service. In many ways ImageJ
} support is better than the best commercial service because you
} work with the person that wrote the code. It is not uncommon for
} someone on the plugin mailing list to help you modify a plugin
} for you needs. There are no commercial vendors that can turn out
} code this way.
}
} I have worked for software companies as a programmer and letting
} the customer work with the programmer is just not done. It uses
} too much of the programmers time and the customer finds out more
} than management wants them to know about the product. Because
} the programmers don't pretend that the patched up out of date
} software that most products are made out of are anything else
} but what it is. Most of the time management doesn't have the
} money to write new software from scratch or look at Cisco when
} they tried to write a new operating system for their routers and
} got in disagreement with the programmers and the programmers
} quit and went into business as Juniper who is now Ciscoes'
} biggest competitor in their high end most profitable business.
}
} Source code is available for most plugins and ImageJ will
} compile them giving you a complete tool for learning to write
} plugins if you want to. Most of use will be content to play the
} sorcerer's apprentice and charge a few lines to make a plugin
} that is almost what we do just what we want and not become full
} blown programmers. But that is very valuable option when you
} need something slightly different than the person that developed
} the plug in needed. A Google search for 'imagej plugin' brings
} up almost 10,000 hits I am sure many of those are duplicates but
} 3,000 to 5,000 plugins is not an unreasonable estimate.
}
} Best Regards
} Gordon
} Gordon Couger
}
} I collect links on information related to light microscopes.
} www.couger.com/microscope/links/gclinks.html
} Please forward anything you think might be useful to others.
} Microscope Documentation is at www.science-info.org
}
} by way of MicroscopyListserver wrote:
} }
} } Email: xiaohutang-at-gmail.com
} } Name: Xiaohu Tang
} }
} } Organization: TU Delft, the Netherlands
} }
} } Title-Subject: [Microscopy] [Filtered] 3D image
} reconstruction software for stereo SEM
} }
} } Question: Hello everyone,
} }
} } I am looking for a 3D image reconstruction software for
} stereo SEM to measure the sample volume. It looks like the MeX
} from Alicona is the only choice so far but the price is too
} expensive. Could anybody give me an idea for a cheaper software
} or a better method to measure the volume by SEM? Thank you.
} }
} } Xiaohu Tang
} } Microlab
} } CiTG of Delft Technology University
} } The Netherlands
} } x.tang-at-citg.tudelft.nl
} }
}
}



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 12:56:11 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Wed, 22 Jun 2005 10:55:24 -0700
Subject: [Microscopy] TEM: Cone angle of LaB6 emitter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Joe;
I have used 60 degree cone emitters with great success on a
2000FX (don't recall the tip radius) and 60 degree cones with 5 micron
radius on a Topcon 002B. They add a nice bit of crispness to the high
resolution image, although they are no substitute for field emission. As
for brightness; no problem. As for lifetime, they don't last quite as
long at a 90 degree, so if you are budget constrained that might be an
issue, but for us it was well worth the minor extra operational cost.
Six months was typical, for a TEM used a lot every day. Usually the
wehnelt needed cleaning once or twice during the life of the tip so
maintenance effort was about the same as a 90 degree. Unfortunately,
Denka, which was our preferred source for that tip on the Topcon, has
stopped manufacturing that particular source.

John Mardinly

This opinion is the opinion of the auther and does not represent the
opinion of Intel Corporation.

-----Original Message-----
} From: Joe Kulik [mailto:juk12-at-psu.edu]
Sent: Tuesday, June 21, 2005 7:10 PM
To: Microscopy-at-microscopy.com

On our JEM-2010 we use a LaB6 emitter with a 90 degree cone and 15
micron tip.

We are considering using a smaller cone angle and/or smaller tip radius.

Does anyone have any comments concerning whether the supposed
improvement in brightness and coherence is worth the shortened lifetime?

_________________________________

Joseph Kulik
Research Associate
Materials Research Institute
The Pennsylvania State University
194 MRI Bldg
University Park, PA 16802

Telephone: 814-865-0344
Fax: 814-863-8561
e-mail: juk12-at-psu.edu
_________________________________






From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 13:33:04 2005



From: Geoff McAuliffe :      mcauliff-at-umdnj.edu
Date: Wed, 22 Jun 2005 14:53:33 -0400
Subject: [Microscopy] Re: viaWWW: Confocal microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

I do not know just how much you need to lower the magnification but the
following may just help?

Raise the eucentric stage to its highest position (moving into focus when
the objective lens is weakened) which will reduce the magnification factor
of the objective lens. Possibly lowering the magnification by 10% or more?

Steve Chapman
Senior Consultant Protrain
For electron microscopy consultancy and training world wide
Tel +44 1280 816512 Fax +44 1280 814007
Mobile +44 7802 966067 Web www.emcourses.comood



----- Original Message -----
} From: "Jay Campbell" {microtomy-at-gmail.com}
To: {Microscopy-at-microscopy.com}
Sent: Wednesday, June 22, 2005 12:36 AM

Before you choose an instrument I think you need to reconsider your "no
dedicated technician" approach. A complex piece of multi-user equipment
with no one person to supervise or police its use is a recipe for
diaster. It won't be much good for teaching or research if it is not
working properly. While you may have a service contract, service
engineers are usually spread to thinly to drop by every week for minor
adjustments that no one understands or has time to learn.

Geoff

--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************


by way of MicroscopyListserver wrote:

} Email: gardnere-at-wpunj.edu
} Name: Eileen Gardner
}
} Organization: William Paterson University
}
} Title-Subject: [Microscopy] [Filtered] MListserver: Confocal microscope
}
} Question: I am the chair of a Biology department at a state university in NJ. I am planning on purchasing a confocal microscope which would be used for both teaching and research. Do you have any ideas as to which brands would be easiest to use and maintain? There will be no technician dedicated to this microscope. Thanks.
}
} ---------------------------------------------------------------------------
}
}




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 14:24:36 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Wed, 22 Jun 2005 12:24:32 -0700
Subject: [Microscopy] Re: TEM digitisation: digital plates vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Kerrie
Digital plate story is discussed nearly every year at ListServer. I don't
want to repeat everything what was discussed (perhaps you could check
archives). From my prospective, digital plates has only one positive side -
the linearity is superior. It may be useful if you do a lot of diffraction
work. For biological samples if you satisfied with film's linearity, then
why you would use image (digital) plates? From practical point of view,
image plates is quite inconvenient: you need to use nearly identical
procedure of loading, unloading, pumping etc as for the film and you are
limited in the number of images you could produce (same as film - 50). You
have to load plates manually in the scanner. Most procedures should be
performed in the dark I believe. With digital camera you have less
resolution but you could generate practically unlimited amount of images (I
normally produce about 100 images in 40 min) and vacuum in the microscope
is great (because you don't have to vent the camera). Images available
immediately (you don't need to wait for scanning - few min per plate I
believe). So, I don't think, image plates much suited for biological
applications keeping in mind the price etc. A few words about
resolution. Somebody on ListServer calculate that single EM film contains
about 7 Gbyte of information - it's enormous amount of information, which
we never use 100%. We normally cropped our images and print them small
(look on the Science magazine pages!), so we dramatically reduce the
resolution on the final "product". Then, such "final product" becomes
comparable with what we have with digital cameras. Keep in mind that even
super-duper printers produce gray tones with resolution about 150 dpi (may
be 300 at the best). Personally, I was surprised to see that 3x3" images
from my BioScan 1Mpix camera printed by good inkjet printer with "gray
inks" practically undistinguishable for the real photo-print. The
disadvantage of such low-res images - you could not crop or enlarge them,
so you just took many pictures at different magnification. For big field
you could combine a few pictures quite easily. I hope it'll help. Have a
great day, Sergey

P.S. A new argument against image plates just cames to me: laser. They
used, I guess, similar laser as for phosphorimager. That laser has limited
life - something like 3 years and should be replaced then. Laser itself is
about $20K, so you need to keep expensive service contract or be ready to
invest in a new laser 2 or 3 year later... I think, CCD chip and phosphor
screen on digital cameras deteriorated as well- mine is still OK after 3
years of extensive use.

At 06:52 AM 6/20/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 15:58:01 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Wed, 22 Jun 2005 13:56:59 -0700
Subject: [Microscopy] Re: Re: viaWWW: 3D image reconstruction software for stereo SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Jun 22, 2005, at 9:50 AM, Xiaohu Tang wrote:

} Gordon, thank you for the valuable information. I checked ImageJ and
} could not find the proper plugin for my problem: I want to use the SEM
} stereo image pairs by tilting the sample and find out the volume data
} from the 3D reconstruction. Based on my understanding (I am not so
} familar with this area), most of the ImageJ plugins about 3D volume
} depend on a series of image slices but not like my case. All the
} suggestions and information will be highly appreciated.
}
Dear Xiaohu.
If Sterecon is still out there, it may suit your needs. It was used
in Albany NY some years ago, and some of the people there may know
whether it is still available.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 16:24:07 2005



From: Rick A. Harris :      raharris-at-ucdavis.edu
Date: Wed, 22 Jun 2005 14:20:38 -0700
Subject: [Microscopy] Re: viaWWW: Confocal microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have to agree with Geoff and others on this list. I have been in this
business for over 25 years and I have never seen a situation where a
piece of equipment with the complexity of a confocal or an SEM or a TEM
or FIB or STM or you name it can be maintained at factory spec without a
dedicated support person to manage the device and train users. Over and
over again at this campus I have seen PIs win grants for equipment
without first having thought through the idea of dedicated support.
They will assume an overloaded tech or lab maanger can do it in their
"spare time". That won't work and eventually the machine will not be
capable of quality work or reliability. The equipement wll be used less
and less and within a short time will be salvaged out. Then, in a year
or three, another PI will need that same piece of equipment and we will
start the process again.

Rick Harris
Unniversity of California
Davis, CA



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 16:53:03 2005



From: Mike Bode :      Mike.Bode-at-soft-imaging.net
Date: Wed, 22 Jun 2005 15:51:04 -0600
Subject: [Microscopy] viaWWW: 3D image reconstruction software for stereo

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Hello Xiaohu,

Stereo techniques usually give you a surface reconstruction, not a volume
reconstruction. Whether you can get meaningful volumetric data depends on
what you are looking at. For example: consider a spherical object lying on a
surface. Stereo images are usually taken at a few degrees tilt apart, so the
two images essentially see the same half of the sphere. You can measure half
of the sphere's volume, but since you cannot see the "backside", you have to
make assumptions. For a sphere, it is fairly easy, but if your object is
irregular, you may not be able to measure its volume without making
assumptions what the "hidden" part looks like.

I am sure you have considered this. I would also point you to our website,
which has information about the stereo module in our software:
http://www.soft-imaging.com/rd/english/405.htm

mike



Michael Bode, Ph.D.
Soft Imaging System Corp.
12596 West Bayaud Avenue
Suite 300
Lakewood, CO 80228
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



-----Original Message-----
} From: by way of MicroscopyListserver [mailto:xiaohutang-at-gmail.com]
Sent: Wednesday, June 22, 2005 07:05
To: microscopy-at-microscopy.com

Below is the result of your feedback form (NJZFM-ultra-55). It was
submitted by (xiaohutang-at-gmail.com) from
http://www.msa.microscopy.org/MicroscopyListserver/MLFormMail.html on
Wednesday, June 22, 2005 at 02:27:04
---------------------------------------------------------------------------

Email: xiaohutang-at-gmail.com
Name: Xiaohu Tang

Organization: TU Delft, the Netherlands

Title-Subject: [Microscopy] [Filtered] 3D image reconstruction software for
stereo SEM

Question: Hello everyone,

I am looking for a 3D image reconstruction software for stereo SEM to
measure the sample volume. It looks like the MeX from Alicona is the only
choice so far but the price is too expensive. Could anybody give me an idea
for a cheaper software or a better method to measure the volume by SEM?
Thank you.

Xiaohu Tang
Microlab
CiTG of Delft Technology University
The Netherlands
x.tang-at-citg.tudelft.nl

---------------------------------------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 16:57:45 2005



From: Ken Converse :      kenconverse-at-qualityimages.biz
Date: Wed, 22 Jun 2005 20:07:00 -0400
Subject: [Microscopy] Re: viaWWW: Confocal microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


-----Original Message-----
} From: "Rick A. Harris" {raharris-at-ucdavis.edu}

The perspective from the Field Service Engineer:

If I had to drop by every customer every week or so for minor "stuff", I
would be: 1) out of business (broke), 2) out of my mind 3) out of customers
because I can't visit all of them that often and the ones who had REAL
problems expect (rightly so) to be taken care of quickly and thoroughly.

PLEASE, PLEASE, PLEASE do not expect your field service engineer to
substitute for a good technician. It is unfair to ALL involved.

If you want your service organization to provide full-time service, expect
to pay 10 or 15 times what you're currently paying for your service
contract.

Ken Converse

owner
QUALITY IMAGES
Servicing Scanning Electron Microscopes
Since 1981
16 Creek Rd.
Delta, PA 17314
717-456-5491
Fax 717-456-7996
kenconverse-at-qualityimages.biz
qualityimages.biz

-----Original Message-----
} From: Geoff McAuliffe [mailto:mcauliff-at-umdnj.edu]
Sent: Wednesday, June 22, 2005 2:54 PM
To: by way of MicroscopyListserver
Cc: microscopy-at-microscopy.com

Before you choose an instrument I think you need to reconsider your "no
dedicated technician" approach. A complex piece of multi-user equipment
with no one person to supervise or police its use is a recipe for
diaster. It won't be much good for teaching or research if it is not
working properly. While you may have a service contract, service
engineers are usually spread to thinly to drop by every week for minor
adjustments that no one understands or has time to learn.

Geoff

--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************


by way of MicroscopyListserver wrote:

} Email: gardnere-at-wpunj.edu
} Name: Eileen Gardner
}
} Organization: William Paterson University
}
} Title-Subject: [Microscopy] [Filtered] MListserver: Confocal microscope
}
} Question: I am the chair of a Biology department at a state university in
NJ. I am planning on purchasing a confocal microscope which would be used
for both teaching and research. Do you have any ideas as to which brands
would be easiest to use and maintain? There will be no technician dedicated
to this microscope. Thanks.
}
} ---------------------------------------------------------------------------
}
}






___________________________________________________________
$0 Web Hosting with up to 200MB web space, 1000 MB Transfer
10 Personalized POP and Web E-mail Accounts, and much more.
Signup at www.doteasy.com




From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 21:27:40 2005



From: Qi Zhang :      qizhang-at-physics.unc.edu
Date: Wed, 22 Jun 2005 22:26:30 -0400
Subject: [Microscopy] JEM-2010F desiccator

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Colleagues,

We have been having oil leak problem with our JEM-2010F desiccator. We
have had 3 desiccators (EMDSC-U10A) and had leaking every 5-6 months. We
were told that the parts replacement wouldn't help so much, because this
model had design problem with about 40% failure possibility. Now, we are
looking for new reliable type of desiccator. Does anyone have any
suggestion? Thanks in advance!

With best regards,
********************************************************
Qi Zhang, Ph.D
164 Phillips Hall, CB#3255
Department of Physics and Astronomy
the University of North Carolina at Chapel Hill
Chapel Hill, NC 27599
Tel: 919-843-2407 (Office)
919-843-0974 (EM Lab)
Fax: 919-962-0480
E-mail: qizhang-at-physics.unc.edu



From MicroscopyL-request-at-ns.microscopy.com Wed Jun 22 22:12:14 2005



From: Matt Olszta :      mjo10-at-psu.edu
Date: Wed, 22 Jun 2005 23:09:56 -0400
Subject: [Microscopy] TEM-sample prep

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

All,
I am currently trying to make TEM samples out of very porous sintered
capacitor materials (Nb and NbO) in which I need to collect stoichiometric
data via EELS analysis. The problem in making these samples is that they
do not polish or ion mill well due to their porous morphology. I had tried
to infiltrate with M-bond, which helped considerably, but even after plasma
treating the samples before analysis, a large C-K edge remained throughout
the sample (which unfortunately sits right in between the elements of
interest). I was wondering if anyone had any suggestions of a filler that
could be used to infiltrate a porous body and then could be easily removed
post ion thinning. Any help would be greatly appreciated.

Regards
Matt Olszta Ph.D.
Materials Research Institute
Penn State University



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 01:02:01 2005



From: :      Ross.Hamilton-at-csl.com.au
Date: Thu, 23 Jun 2005 15:03:45 +1000
Subject: [Microscopy] TEM Negative Stains for Vaccine Formulations

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

We wish to negatively stain some vaccine formulations.
Does anyone know of a negative stain that you can use at acid pH that is not a uranyl salt?
Does anyone know of a negative stain that you can use at basic pH that is not sodium tetraborate? We have used sodium tetraborate but it tends to "bubble" under the beam!

Ross Hamilton,
Senior Scientist,
E.M. Unit,
Bioanalytical Sciences,
CSL Limited,
45 Poplar Road,
Parkville,3052
Victoria,Australia.
+61 3 9389 1307



****************************************************************************************************************
This email and any attached files are intended solely for the named addressee, are confidential and may contain legally privileged information. The copying or distribution of them or any information they contain, by anyone other than the addressee, is prohibited. If you have received this email in error, please let us know by telephone or return the email to the sender
and destroy all copies. Thank you.

CSL Limited A.C.N. 051 588 348
45 Poplar Road Parkville Victoria 3052 Australia
Phone: +61 3 9389 1911 Fax: +61 3 9389 1434
***************************************************************************************************************




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 07:35:37 2005



From: Bill Miller :      microbill-at-mohawk.net
Date: Thu, 23 Jun 2005 08:34:55 -0400
Subject: [Microscopy] Re: Re: TEM digitisation: digital plates vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sounds sort of like Pixel envy doesn't it? - since the new imaging
plate systems can make images with up to 110Mega pixels - you'd need
to shoot more than 100 pictures at 1Meg each to get equivalent
images - and by the way the systems use diode lasers that have very
long life and they don't cost anywhere near $20K


Imaging plates are not necessarily the right solution for every
situation but to categorically state that they are useless for
biological use is simply wrong. Like with any system there are
strengths and weaknesses but if nothing else the full field of view
pixels/$ is unbeatable.


Bill Miller

At 03:24 PM 6/22/2005, Sergey Ryazantsev wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 08:35:45 2005



From: Nestor J. Zaluzec :      zaluzec-at-microscopy.com
Date: Thu, 23 Jun 2005 08:35:03 -0500
Subject: [Microscopy] EPON-what is is made of?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Colleagues...

Can someone tell me the nominal consitutents of EPON used
for embedding TEM biological samples. Specifically what
are it's major elemental components. (C, H, O, N, ......???)

It is not obvious to me which of the many EPONS listed in a Google search is
the one typically used by the life science community in
preparing TEM sections.



Thanks in advance,

Nestor
Your Friendly Neighborhood SysOp


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 08:35:51 2005



From: Tom Phillips :      phillipst-at-missouri.edu
Date: Thu, 23 Jun 2005 08:35:03 -0500
Subject: [Microscopy] Re: TEM Negative Stains for Vaccine Formulations

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

check out www.nanoprobes.com they sell tungstate and vanadate based
negative stains. i have never used them so can't comment.

At 12:03 AM 06/23/05, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thomas E. Phillips, PhD
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400

573-882-4712 (office)
573-882-0123 (fax)
PhillipsT-at-missouri.edu




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 08:45:07 2005



From: dwaugh-at-kent.edu (by way of MicroscopyListserver)
Date: Thu, 23 Jun 2005 08:44:21 -0500
Subject: [Microscopy] viaWWW: film for contact microradiographs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (dwaugh-at-kent.edu) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Thursday, June 23, 2005 at 08:34:28
---------------------------------------------------------------------------

Email: dwaugh-at-kent.edu
Name: David Waugh

Organization: Kent State University, Geology

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: I'm looking to make contact microradiographs of crab claw thin sections (~40 microns thick). We have an x-ray machine, the problem is the film. Doing a google search I have found two Kodak films that are used, but can't figure out if they are still in production. Kodak SO-343, seems to be popular, I can't tell if it is an xray or holographic film, but either way there is no mention of it on the Kodak website or from any supplier, same thing for Kodak "spectroscopic film". A film "Fuji B&W POS/71337" that I found in one paper, does not seem to exist. I need something that is very high resolution, and preferable can be developed with standard B&W chemistry. Any ideas on what the current films in use are? Would a conventional film like Kodak T-max work (we have that film and chemical already lying around in the now mostly mothballed darkroom)? Thanks, David

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 08:55:09 2005



From: Mike Nesta :      mnesta-at-ebsciences.com
Date: Thu, 23 Jun 2005 09:54:27 -0400
Subject: [Microscopy] Re: viaWWW: Siemens IA Filaments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

At Energy Beam Sciences we can rebuild any tungsten filament, including
Siemens. We also provide new mounted filaments for most Microscopes. You
can contact us on (800) 992-9037 or by email at ebs-at-ebsciences.com.

Michael R. Nesta
Managing Director
Energy Beam Sciences, Inc.
Tel: 860 653-0411
Fax: 860 653-0422
MNesta-at-ebsciences.com
www.ebsciences.com
“ADDING BRILLIANCE TO YOUR VISION”



by way of MicroscopyListserver wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 09:08:42 2005



From: twigg-at-estd.nrl.navy.mil
Date: Thu, 23 Jun 2005 10:09:16 -0400
Subject: [Microscopy] Beam stop and TEM diffraction pattern recording by digital camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am posed to finally make the leap to a digital camera (a Gatan
Ultrascan 1000) after years of SO-163. One potential problem that I
face, however, is digital recording of SAD diffraction patterns. My
top-entry Hitachi H-9000UHR HRTEM did not come with a beam stop (that
rod or pointer thing that you use to block the main beam while
recording a diffraction pattern) and it is therefore my concern that
I might not be able to record diffraction patterns without damaging
the digital camera. Hitachi does not seam to have a beam stop thing
for my vintage (1991) H-9000, so I am behind the eight ball in this
regard. I hear from Gatan that I should try to avoid using film once
the Ultrascan is installed, so that that flaked-off film or whatever
does not fall onto the digital camera and mess it up, so using film
to record diffraction patterns may not be an option either. Any
suggestions?

Thanks,

Mark
--
Dr. Mark E. Twigg
Code 6812
Naval Research Laboratory
4555 Overlook Ave., S.W.
Washington DC 20375

tel: (202) 404-8543
fax: (202) 404-7194


From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 09:18:16 2005



From: Tom Phillips :      phillipst-at-missouri.edu
Date: Thu, 23 Jun 2005 09:17:27 -0500
Subject: [Microscopy] Re: EPON-what is is made of?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html



Epon 812 was a Shell patented formulation and has now been replaced by
numerous "identical chemical equivalents" according to EM suppliers but
rarely are the formulations given. It is a

glycerol polyglycidyl ether (i think it is a mixture of different glycidyl
ethers).

"Epon resins" in lab formulations usually also contain NMA (Nadic©Methyl
Anhydride
(Methyl-5-Norbornene-2,3-Dicarboxylic Anhydride) as a hardner

C10H10O3

and

DDSA (Dodecenyl Succinic Anhydride) (about 2:1:1 for epon/nma/ddsa but this
varies in different labs) plus either DMP-30
Tri-(dimethylaminomethyl) phenol

C15H29N8OH or BDMA as catalysts

Tom


At 08:35 AM 06/23/05, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thomas E. Phillips, PhD
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400

573-882-4712 (office)
573-882-0123 (fax)
PhillipsT-at-missouri.edu





From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 09:20:25 2005



From: Mike Nesta :      mnesta-at-ebsciences.com
Date: Thu, 23 Jun 2005 10:19:39 -0400
Subject: [Microscopy] Re: RE: TEM: Cone angle of LaB6 emitter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear John,

The discontinued Denka source you are referring to was the M-7, which
featured a hard mounted tip. All of the tip configurations that were
available on the M-7 are still available on the M-3, wire mounted tip,
including the 60 degree cone angle/5 micron radius configuration you
referred to. The reason Denka discontinued the M-7 was that they
discovered through internal testing that the M-3 provides stability
equal to or greater than the M-7. Denka feels that this discovery
rendered the higher cost for the M-7 an unnecessary expense.

Disclaimer - Energy beam Sciences, Inc. is the US distributor for Denka
LaB6 and Thermal Field Emission Cathodes.

Michael R. Nesta
Managing Director
Energy Beam Sciences, Inc.
Tel: 860 653-0411
Fax: 860 653-0422
MNesta-at-ebsciences.com
www.ebsciences.com
“ADDING BRILLIANCE TO YOUR VISION”





Mardinly, John wrote:

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 10:00:02 2005



From: Sinkler, Wharton :      Wharton.Sinkler-at-uop.com
Date: Thu, 23 Jun 2005 09:59:24 -0500
Subject: [Microscopy] Beam stop and TEM diffraction pattern recording by

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


Mark,

I routinely use our US-1000 for single crystal diffraction patterns with
no beam stop. This is because the central beam is actually fairly weak
in this situation, as so much intensity is getting diffracted. If this
is the large majority of your diffraction work you will be fine.

On the other hand, when collecting polycrystalline ring patterns or
patterns from amorphous samples, you will have a problem. The rings are
typically weak and the central spot is strong. To get decent signal to
noise in the rings you need to up the exposure time or dose. This is
where I use the beam stop so if you do a lot of this you should consider
making one. Without the stop you'll pretty heavily saturate the CCD,
producing a large streak across the central spot.

Talk to Gatan about conditions which would damage the phosphor on the
CCD, but I think they are rather extreme - typically these are pretty
robust for normal TEM and diffraction conditions.

Regards,
Wharton

*************************************************************
Wharton Sinkler, PhD.
UOP LLC
50 E. Algonquin Rd.
Des Plaines IL 60017-5017
mailto: wharton.sinkler-at-uop.com
tel 847-391-3878
fax 847-391-3719



-----Original Message-----
} From: twigg-at-estd.nrl.navy.mil [mailto:twigg-at-estd.nrl.navy.mil]
Sent: Thursday, June 23, 2005 9:09 AM
To: Microscopy-at-microscopy.com

I am posed to finally make the leap to a digital camera (a Gatan
Ultrascan 1000) after years of SO-163. One potential problem that I
face, however, is digital recording of SAD diffraction patterns. My
top-entry Hitachi H-9000UHR HRTEM did not come with a beam stop (that
rod or pointer thing that you use to block the main beam while
recording a diffraction pattern) and it is therefore my concern that
I might not be able to record diffraction patterns without damaging
the digital camera. Hitachi does not seam to have a beam stop thing
for my vintage (1991) H-9000, so I am behind the eight ball in this
regard. I hear from Gatan that I should try to avoid using film once
the Ultrascan is installed, so that that flaked-off film or whatever
does not fall onto the digital camera and mess it up, so using film
to record diffraction patterns may not be an option either. Any
suggestions?

Thanks,

Mark
--
Dr. Mark E. Twigg
Code 6812
Naval Research Laboratory
4555 Overlook Ave., S.W.
Washington DC 20375

tel: (202) 404-8543
fax: (202) 404-7194




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 10:18:22 2005



From: Technik :      technik-at-ditabis.de
Date: Thu, 23 Jun 2005 17:18:32 +0200
Subject: [Microscopy] Re: TEM digitisation: digital plates vs digital camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Community ,

I think the idea of the list server is to share experience one has made
in any topic
with the microscopy community. But this assumes that one has made this
experience personally.
Probably everybody believes and guesses a lot about instruments,
companies and persons but shall such
guesswork be shared on such a public experience based forum like this?

Please let me point out what up-to-date laser technology means:
Sergey is right, former laser had a limited life time. Such gas lasers
being produced some decades ago had this limitations
whereas modern semiconductor lasers as used in the DITABIS MICRON are
very stable and have a life time comparable to CCD sensors.
I am disappointed to see it obviously has become more and more common to
read "personal opinions" of some forum members
not being based on gained knowledge.

The resolution discussion: why does anybody want to have high resolution
systems when downsizing the
resolution and more important loosing the information which should be
the aim of any scientifical work.
I know that our scientific customers do need the information and don't
make high resolution images just for their personal satisfaction.

Another point mentioned by John Mardinly was that the imaging plate
store only electrons.
In principle he is right except that you can detect nearly every high
energy radiation with imaging plates.
With former imaging plate systems an identification of the imaging plate
- and thus the exposure - was impossible.
The Ditabis imaging plate system is able to optically scan the surface
(e.g. labeled with unique numbers.....) in parallel
to normal image scans. That label information is stamped into the final
image in one and the same run.
A software package even allows to merge TEM settings information into
the header of each respective image.

Best regards,
Andreas Elkeries

Dipl.-Ing.
Customer Support / R&D

-------------------------------------
DITABIS AG
Digital Biomedical Imaging Systems AG
Stuttgarter Str. 13
D-75179 Pforzheim, Germany

Tel.: +49 (0)7231 1559893
Fax.: +49 (0)7231 589791

a.elkeries-at-ditabis.de
http://www.ditabis.de
=====================================



} Sergey Ryazantsev wrote:
}
} ------------------------------------------------------------------------------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -------------------------------------------------------------------------------
}
}
} Kerrie
} Digital plate story is discussed nearly every year at ListServer. I
} don't want to repeat everything what was discussed (perhaps you could
} check archives). From my prospective, digital plates has only one
} positive side - the linearity is superior. It may be useful if you do
} a lot of diffraction work. For biological samples if you satisfied
} with film's linearity, then why you would use image (digital) plates?
} From practical point of view, image plates is quite inconvenient: you
} need to use nearly identical procedure of loading, unloading, pumping
} etc as for the film and you are limited in the number of images you
} could produce (same as film - 50). You have to load plates manually in
} the scanner. Most procedures should be performed in the dark I
} believe. With digital camera you have less resolution but you could
} generate practically unlimited amount of images (I normally produce
} about 100 images in 40 min) and vacuum in the microscope is great
} (because you don't have to vent the camera). Images available
} immediately (you don't need to wait for scanning - few min per plate I
} believe). So, I don't think, image plates much suited for biological
} applications keeping in mind the price etc. A few words about
} resolution. Somebody on ListServer calculate that single EM film
} contains about 7 Gbyte of information - it's enormous amount of
} information, which we never use 100%. We normally cropped our images
} and print them small (look on the Science magazine pages!), so we
} dramatically reduce the resolution on the final "product". Then, such
} "final product" becomes comparable with what we have with digital
} cameras. Keep in mind that even super-duper printers produce gray
} tones with resolution about 150 dpi (may be 300 at the best).
} Personally, I was surprised to see that 3x3" images from my BioScan
} 1Mpix camera printed by good inkjet printer with "gray inks"
} practically undistinguishable for the real photo-print. The
} disadvantage of such low-res images - you could not crop or enlarge
} them, so you just took many pictures at different magnification. For
} big field you could combine a few pictures quite easily. I hope it'll
} help. Have a great day, Sergey
}
} P.S. A new argument against image plates just cames to me: laser.
} They used, I guess, similar laser as for phosphorimager. That laser
} has limited life - something like 3 years and should be replaced
} then. Laser itself is about $20K, so you need to keep expensive
} service contract or be ready to invest in a new laser 2 or 3 year
} later... I think, CCD chip and phosphor screen on digital cameras
} deteriorated as well- mine is still OK after 3 years of extensive use.
}
} At 06:52 AM 6/20/2005, you wrote:
}
}
} } ------------------------------------------------------------------------------
} }
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe --
} } http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} } -------------------------------------------------------------------------------
} }
} }
} } I have been given the unenviable task of having to decide between a
} } digital
} } camera system and a digital imaging plate system (the Ditabis system) to
} } enable our older TEMs to become fully digital imaging systems. We
} } have Jeol
} } 1200EX and 100CX instruments and a Phillips CM10. Has anyone had
} } experience
} } of the Ditabis system? In the UK there is only one currently in use,
} } and I
} } have contact with that person. Our samples are biological samples, some
} } routine clinical specimens, but also research samples and cell culture
} } samples. We currently use Kodak em film 4489 and require the
} } replacement
} } system to have comparable resolution.
} } Any comments would be gratefully received.
}
}
} _____________________________________
}
} Sergey Ryazantsev Ph. D.
} Electron Microscopy
} UCLA School of Medicine
} Department of Biological Chemistry
} 10833 Le Conte Ave, Room 33-080
} Los Angeles, CA 90095
}
} Phone: (310) 825-1144 (office)
} (310) 206-1029 (Lab)
} FAX (departmental): (310) 206-5272
} mailto:sryazant-at-ucla.edu
}
}
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 11:11:34 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Thu, 23 Jun 2005 09:10:46 -0700
Subject: [Microscopy] Re: RE: TEM: Cone angle of LaB6 emitter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Michael;
Thank you for this good news. When I attempted to order an M7 a
few months ago from another vendor of Denka emitters, that information
was not available.

John Mardinly

-----Original Message-----
} From: Mike Nesta [mailto:mnesta-at-ebsciences.com]
Sent: Thursday, June 23, 2005 7:20 AM
To: Mardinly, John
Cc: Joe Kulik; Microscopy-at-microscopy.com




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 11:18:08 2005



From: :      nyilmaz-at-mersin.edu.tr
Date: Thu, 23 Jun 2005 19:17:26 +0300
Subject: [Microscopy] bacteria and bone marrow processing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello Everybody

We're planning to study bacteria suspension (Staphylococcus aureus) and bone marrow with TEM. Does anybody help us for bacteria and bone marrow (about 1 ml) processing protocol for TEM?

Thanks...

Dr. Necat Yilmaz








From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 11:34:31 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Thu, 23 Jun 2005 09:33:16 -0700
Subject: [Microscopy] Re: TEM digitisation: digital plates vs digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Andreas Elkeries;
As Nestor has pointed out, it is inappropriate to make
disparaging remarks about individuals in the MSA Listserver, especially
as to whether their comments come from personal experience or
hallucination. It is especially surprising to see such obstreperousness
coming from a member of any 'Customer Support' staff. In my particular
case, I conducted an evaluation of Ditabis imaging plates through an
authorized US dealer of your product, as Mr. Miller can attest to. The
full results of that evaluation cannot be shared on the Listserver, but
I can say that no microscope data that would normally be printed on film
was present on the images I received, and in discussions with Mr.
Miller, there was never communicated any expectation that the microscope
data could ever be imprinted on imaging plates. If there is a new
development that would allow that, I think we would all be interested in
hearing more details about it.

John Mardinly


-----Original Message-----
} From: Technik [mailto:technik-at-ditabis.de]
Sent: Thursday, June 23, 2005 8:19 AM
To: Microscopy-at-microscopy.com; sryazant-at-ucla.edu

Dear Community ,

I think the idea of the list server is to share experience one has made
in any topic
with the microscopy community. But this assumes that one has made this
experience personally.
Probably everybody believes and guesses a lot about instruments,
companies and persons but shall such
guesswork be shared on such a public experience based forum like this?

Please let me point out what up-to-date laser technology means:
Sergey is right, former laser had a limited life time. Such gas lasers
being produced some decades ago had this limitations
whereas modern semiconductor lasers as used in the DITABIS MICRON are
very stable and have a life time comparable to CCD sensors.
I am disappointed to see it obviously has become more and more common to

read "personal opinions" of some forum members
not being based on gained knowledge.

The resolution discussion: why does anybody want to have high resolution

systems when downsizing the
resolution and more important loosing the information which should be
the aim of any scientifical work.
I know that our scientific customers do need the information and don't
make high resolution images just for their personal satisfaction.

Another point mentioned by John Mardinly was that the imaging plate
store only electrons.
In principle he is right except that you can detect nearly every high
energy radiation with imaging plates.
With former imaging plate systems an identification of the imaging plate

- and thus the exposure - was impossible.
The Ditabis imaging plate system is able to optically scan the surface
(e.g. labeled with unique numbers.....) in parallel
to normal image scans. That label information is stamped into the final
image in one and the same run.
A software package even allows to merge TEM settings information into
the header of each respective image.

Best regards,
Andreas Elkeries

Dipl.-Ing.
Customer Support / R&D

-------------------------------------
DITABIS AG
Digital Biomedical Imaging Systems AG
Stuttgarter Str. 13
D-75179 Pforzheim, Germany

Tel.: +49 (0)7231 1559893
Fax.: +49 (0)7231 589791

a.elkeries-at-ditabis.de
http://www.ditabis.de
=====================================



} Sergey Ryazantsev wrote:
}
}
------------------------------------------------------------------------
------
}
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
}
------------------------------------------------------------------------
-------
}
}
} Kerrie
} Digital plate story is discussed nearly every year at ListServer. I
} don't want to repeat everything what was discussed (perhaps you could
} check archives). From my prospective, digital plates has only one
} positive side - the linearity is superior. It may be useful if you do
} a lot of diffraction work. For biological samples if you satisfied
} with film's linearity, then why you would use image (digital) plates?
} From practical point of view, image plates is quite inconvenient: you
} need to use nearly identical procedure of loading, unloading, pumping
} etc as for the film and you are limited in the number of images you
} could produce (same as film - 50). You have to load plates manually in

} the scanner. Most procedures should be performed in the dark I
} believe. With digital camera you have less resolution but you could
} generate practically unlimited amount of images (I normally produce
} about 100 images in 40 min) and vacuum in the microscope is great
} (because you don't have to vent the camera). Images available
} immediately (you don't need to wait for scanning - few min per plate I

} believe). So, I don't think, image plates much suited for biological
} applications keeping in mind the price etc. A few words about
} resolution. Somebody on ListServer calculate that single EM film
} contains about 7 Gbyte of information - it's enormous amount of
} information, which we never use 100%. We normally cropped our images
} and print them small (look on the Science magazine pages!), so we
} dramatically reduce the resolution on the final "product". Then, such

} "final product" becomes comparable with what we have with digital
} cameras. Keep in mind that even super-duper printers produce gray
} tones with resolution about 150 dpi (may be 300 at the best).
} Personally, I was surprised to see that 3x3" images from my BioScan
} 1Mpix camera printed by good inkjet printer with "gray inks"
} practically undistinguishable for the real photo-print. The
} disadvantage of such low-res images - you could not crop or enlarge
} them, so you just took many pictures at different magnification. For
} big field you could combine a few pictures quite easily. I hope it'll

} help. Have a great day, Sergey
}
} P.S. A new argument against image plates just cames to me: laser.
} They used, I guess, similar laser as for phosphorimager. That laser
} has limited life - something like 3 years and should be replaced
} then. Laser itself is about $20K, so you need to keep expensive
} service contract or be ready to invest in a new laser 2 or 3 year
} later... I think, CCD chip and phosphor screen on digital cameras
} deteriorated as well- mine is still OK after 3 years of extensive use.
}
} At 06:52 AM 6/20/2005, you wrote:
}
}
} }
------------------------------------------------------------------------
------
} }
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} } To Subscribe/Unsubscribe --
} } http://www.microscopy.com/MicroscopyListserver
} } On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} }
------------------------------------------------------------------------
-------
} }
} }
} } I have been given the unenviable task of having to decide between a
} } digital
} } camera system and a digital imaging plate system (the Ditabis system)
to
} } enable our older TEMs to become fully digital imaging systems. We
} } have Jeol
} } 1200EX and 100CX instruments and a Phillips CM10. Has anyone had
} } experience
} } of the Ditabis system? In the UK there is only one currently in use,

} } and I
} } have contact with that person. Our samples are biological samples,
some
} } routine clinical specimens, but also research samples and cell
culture
} } samples. We currently use Kodak em film 4489 and require the
} } replacement
} } system to have comparable resolution.
} } Any comments would be gratefully received.
}
}
} _____________________________________
}
} Sergey Ryazantsev Ph. D.
} Electron Microscopy
} UCLA School of Medicine
} Department of Biological Chemistry
} 10833 Le Conte Ave, Room 33-080
} Los Angeles, CA 90095
}
} Phone: (310) 825-1144 (office)
} (310) 206-1029 (Lab)
} FAX (departmental): (310) 206-5272
} mailto:sryazant-at-ucla.edu
}
}
}
}
}





From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 12:25:31 2005



From: Gilles Grondin :      gilles.grondin-at-usherbrooke.ca
Date: Thu, 23 Jun 2005 13:21:34 -0400
Subject: [Microscopy] Fwd: petit contrat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


}
} From: gilles_grondin-at-uqtr.ca
}
} *************************************************************
} Cutting sections from methacrylate-embedded blocks (non-plastination)
}
} Hi,
} Anyone want to cut some plastic sections for us?
}
} We have a number of tissue blocks embedded for light microscopy in hpma, and
} we would like them sectioned at 1.5 - 2 microns, and mounted on slides. We
} aren't currently set up for cutting plastic sections, and if there is anyone
} with some spare microtome capacity who would like to earn some pocket money
} here's your chance. I would like an estimate of the per-slide cost of cutting
} sections and mounting them on a glass slide. Just guessing, initially we
} would
} like about 100 sections from about 20 different blocks.
}
} As an aside, anyone tried to embed blocks with P40 and then cut them for
} light
} microscopy?
}
} Sincerely,
} Dave Griffiths
}
} David Griffiths
} Section of Anatomy and Pathology
} Norwegian School of Veterinary Science
} P. O. Box 8146 Dep
} N-0033 Oslo
} Norway
} Tel: 22964542/Fax: 22964764
} E-mail: david.griffiths-at-veths.no
} ****************************************************
}
} -------------------------------------------------
} Courriel expédié via https://courriel.uqtr.ca




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 18:56:17 2005



From: Bill Miller :      microbill-at-mohawk.net
Date: Thu, 23 Jun 2005 19:55:36 -0400
Subject: [Microscopy] digital imaging plate tracking

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

As rightly pointed out by John Mardinly, a drawback to the imaging
plates has always been the lack of foolproof data and image
correlation. There is now a solution that for some TEMs will
automatically read the microscope parameters and link it with bar
coded plates which are read during the scan process.

A complete description of the tracking system is available though ElectroImage.

Best Regards - Bill Miller




From MicroscopyL-request-at-ns.microscopy.com Thu Jun 23 20:18:28 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Thu, 23 Jun 2005 18:15:48 -0700
Subject: [Microscopy] Re: Re: TEM digitisation: digital plates vs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Bill
This is very common mistake thinking that more pixels gives you 'more'
resolution. If your sample has, let say 10 nm details and nothing smaller,
I would expect that quite decent 'resolution' on this particular sample
would be if you utilize something like 3 pixels per 10 nm =} 3.3
nm/pixel. You simply don't need more pixels if there is nothing smaller on
your sample. Adding more pixels you will basically add more noise to the
image and that's it. Ultrathin sections of biological material, which I
called 'biological samples' have quite poor resolution because of fixation,
dehydratation plastic embedding and so on. Therefore, you will just spoil
those 110M of pixels... It's like some people expected 3A resolution
(resolution of the scope itself) on 80 nm thick poorly infiltrated
chattered plastic sections of biological material, no way: 100A may be...

Again, even on my 3.3GHz Pentium 5 with 2 Gb 3400 RAM computer, I do see
deterioration in performance if I am working on couple 50Mb pictures, so
you really need super computer to handle your 110 Mb pictures. And I am
not talking about printing those images: on my networked color Xerox Phaser
7700GX with a lot of memory (maximum permitted) it takes about 20-30 sec to
sent 10-20 Mb image (and more than minute to print at 1200 dpi). Your files
will simply kill all our UCLA network! I think, you guys have better
network, probably dedicated to handle such giant files.... So, I am just
reasoning, why from my prospective, there is not much use of image plates
in biological applications. I am glad that Ditabis is using solid state
lasers - it definitely will add some juice to the system. By the way, how
long it takes to scan one plate at maximum resolution?

I do belive that image plate technology has a great potential in material
science applications when huge dynamic range is critical. Have a great
day, Sergey

At 05:34 AM 6/23/2005, you wrote:
} Sounds sort of like Pixel envy doesn't it? - since the new imaging plate
} systems can make images with up to 110Mega pixels - you'd need to shoot
} more than 100 pictures at 1Meg each to get equivalent images - and by the
} way the systems use diode lasers that have very long life and they don't
} cost anywhere near $20K
}
}
} Imaging plates are not necessarily the right solution for every situation
} but to categorically state that they are useless for biological use is
} simply wrong. Like with any system there are strengths and weaknesses but
} if nothing else the full field of view pixels/$ is unbeatable.
}
}
} Bill Miller
}
} At 03:24 PM 6/22/2005, Sergey Ryazantsev wrote:
}
}
} } ------------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From MicroscopyL-request-at-ns.microscopy.com Fri Jun 24 02:44:29 2005



From: Bill Miller :      microbill-at-mohawk.net
Date: Fri, 24 Jun 2005 10:15:05 -0400
Subject: [Microscopy] Re: Re: Re: TEM digitisation: digital plates vs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear All,
.. I might add for those interested in getting data from a Philips EM 400,
410, 420 and 430 that there is a interface card you can install without much
knowledge.
It will get you
High tension value
Magnification / diffraction camera length (dependent on chosen camera: TV,
sheet film, 224x36mm port)
Filmcode
Filmnumber
Exposure time on filmilm
Please see http://www.stefan-diller.com/rem_tem3_en.htm.

For Zeiss EM 10 A, B, C there is the same possibility but without
transmission of filmnumbers (but I might develop such an interface if there
is the need).
Values are only
High tension
Magnification
0.4x Magnifying switch
Please see http://www.stefan-diller.com/rem_tem4_en.htm
Installation should be by service technician.

Values are send via RS 232 interface to the computer. There should be no
problem to incorporate them in any imaging software.

Best regards,
Stefan Diller



----- Original Message -----
} From: "Bill Miller" {microbill-at-mohawk.net}
To: {Microscopy-at-microscopy.com}
Sent: Friday, June 24, 2005 1:55 AM

Hello Mark,

Diffraction beam stop is a standard option of SIA TEM CCD cameras. It is
available separately (without a camera) as well.

Vitaly Feingold
SIA
2773 Heath Lane
Duluth, GA 30096
Ph. 770-232-7785
Fax 770-232-1791
www.sia-cam.com

----- Original Message -----
} From: {twigg-at-estd.nrl.navy.mil}
To: {Microscopy-at-microscopy.com}
Sent: Thursday, June 23, 2005 10:09 AM

Sergey - I would normally ignore this kind of
thing but you have missed the point entirely and
I hope this helps to clarify it. The number of
pixels does not necessarily define the resolution
of an image - in this case it defines the field
of view. The pixels can be as small as 7.5um in
the image plane but what counts is that the image
can be as large as ~11,000 x 10,000 pixels. The
high pixel count allowing you to take MANY fewer
pictures to get the same information; so one
picture contains the information (pixels) of 100
1mega pixel images (if both sensors have 7.5um
pixel in the image plane) - I've responded to
your points individually below..

Bill



------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Bill
This is very common mistake thinking that more
pixels gives you 'more' resolution.

} No one ever said that more pixels gave
more resolution - resolution is dependent upon
the pixel size in the image plane as well as
point spread and MTF of the system - more pixels
give a larger field of view not more resolution.

If your sample has, let say 10 nm details and
nothing smaller, I would expect that quite decent
'resolution' on this particular sample would be
if you utilize something like 3 pixels per 10 nm
=} 3.3 nm/pixel. You simply don't need more
pixels if there is nothing smaller on your sample.

} Resolution is one thing but being able to
actually visualize a structure requires
significant oversampling - how confident would
you be of a structure represented by 3 pixels -
would you ever enlarge the image enough to even
really see 3 pixels? The Nyquist sampling limit
is fine for mathematically determining the
resolution limit of an image but over sampling by
a factor of 5 or 10 will give you enough pixels
to actually see the structure. But resolution is
not the relavent point to all of this , it is field of view.

Adding more pixels you will basically add more
noise to the image and that's it.

} More pixels = more noise – an interesting, but
an obviously flawed concept or there would be NO
market for high end digital cameras of any
sort. It might be true if the pixels were
smaller rather than spread over a larger
area. In this case the idea of more pixels is to
expand the field of view not to increase the
resolution. The pixel size determines the
resolution not the pixel number - you are confusing the two.

Ultrathin sections of biological material, which
I called 'biological samples' have quite poor
resolution because of fixation, dehydratation
plastic embedding and so on. Therefore, you will
just spoil those 110M of pixels... It's like
some people expected 3A resolution (resolution of
the scope itself) on 80 nm thick poorly
infiltrated chattered plastic sections of
biological material, no way: 100A may be...

} So with more pixels you can get the 100A
resolution in an image at lower magnification and
a concomitantly larger field of view - again
fewer pictures to shoot -- seems like simple enough concept

Again, even on my 3.3GHz Pentium 5 with 2 Gb 3400
RAM computer, I do see deterioration in
performance if I am working on couple 50Mb
pictures, so you really need super computer to
handle your 110 Mb pictures. So is it easier and
faster to process over 100 images or a single
image that does not have to be stitches to get
all of the information? And I am not talking
about printing those images: on my networked
color Xerox Phaser 7700GX with a lot of memory
(maximum permitted) it takes about 20-30 sec to
sent 10-20 Mb image (and more than minute to print at 1200 dpi).

} Are the people who are scanning their negatives
doing so at 300dpi for fear thet the file would
otherwise be too big? Absolutely not. People pay
a great deal of money for scanners with 7um
pixels so that they get as much information as
possible form their TEM negatives. Big files are
a fact of life in science - how big is a
confocal image made of 100’s of serial images?

simply kill all our UCLA network!

} So don't use it, but that doesn't mean that
someone else might not "feel" like you do.

I think, you guys have better network, probably
dedicated to handle such giant files.... So, I
am just reasoning, why from my prospective, there
is not much use of image plates in biological applications.

} That’s like asking why everybody doesn't drive
a Ferrari!!! And just what direct knowledge of
the use of plates in biology do you have? Just
like with megapixel CCD cameras, price, not
performance is the bigger barrier. How many
people do you think are stuck with a 1meg small
field of view camera for biological work - forced
to shoot many hundreds of images to get a field
of view comparable in size and resolution to what
they have been getting with film - just because
they couldn't afford the 16meg CCD camera? Would
you have bought the 1meg camera if you could have
afforded a 4 or 16meg one? So, I wonder why there
is not much use of 8k x 8k cameras in biological
applications? Increased noise? Nope, it's the Money!

I am glad that Ditabis is using solid state
lasers - it definitely will add some juice to the
system. By the way, how long it takes to scan one plate at maximum resolution?

} a few unattended minutes - far less time than
it takes to shoot and stitch the comparable 100+
1 meg images required to cover the same field of view.

I do belive that image plate technology has a
great potential in material science applications
when huge dynamic range is critical. Have a great day, Sergey






At 05:34 AM 6/23/2005, you wrote:

Sounds sort of like Pixel envy doesn't it? -
since the new imaging plate systems can make
images with up to 110Mega pixels - you'd need to
shoot more than 100 pictures at 1Meg each to get
equivalent images - and by the way the systems
use diode lasers that have very long life and
they don't cost anywhere near $20K


Imaging plates are not necessarily the right
solution for every situation but to categorically
state that they are useless for biological use is
simply wrong. Like with any system there are
strengths and weaknesses but if nothing else the
full field of view pixels/$ is unbeatable.


Bill Miller

At 03:24 PM 6/22/2005, Sergey Ryazantsev wrote:



------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Kerrie
Digital plate story is discussed nearly every
year at ListServer. I don't want to repeat
everything what was discussed (perhaps you could
check archives). From my prospective, digital
plates has only one positive side - the linearity
is superior. It may be useful if you do a lot of
diffraction work. For biological samples if you
satisfied with film's linearity, then why you
would use image (digital) plates? From practical
point of view, image plates is quite
inconvenient: you need to use nearly identical
procedure of loading, unloading, pumping etc as
for the film and you are limited in the number of
images you could produce (same as film - 50). You
have to load plates manually in the
scanner. Most procedures should be performed in
the dark I believe. With digital camera you have
less resolution but you could generate
practically unlimited amount of images (I
normally produce about 100 images in 40 min) and
vacuum in the microscope is great (because you
don't have to vent the camera). Images available
immediately (you don't need to wait for scanning
- few min per plate I believe). So, I don't
think, image plates much suited for biological
applications keeping in mind the price etc. A few
words about resolution. Somebody on ListServer
calculate that single EM film contains about 7
Gbyte of information - it's enormous amount of
information, which we never use 100%. We
normally cropped our images and print them small
(look on the Science magazine pages!), so we
dramatically reduce the resolution on the final
"product". Then, such "final product" becomes
comparable with what we have with digital
cameras. Keep in mind that even super-duper
printers produce gray tones with resolution about
150 dpi (may be 300 at the best). Personally, I
was surprised to see that 3x3" images from my
BioScan 1Mpix camera printed by good inkjet
printer with "gray inks" practically
undistinguishable for the real photo-print. The
disadvantage of such low-res images - you could
not crop or enlarge them, so you just took many
pictures at different magnification. For big
field you could combine a few pictures quite
easily. I hope it'll help. Have a great day, Sergey

P.S. A new argument against image plates just
cames to me: laser. They used, I guess, similar
laser as for phosphorimager. That laser has
limited life - something like 3 years and should
be replaced then. Laser itself is about $20K, so
you need to keep expensive service contract or be
ready to invest in a new laser 2 or 3 year
later... I think, CCD chip and phosphor screen
on digital cameras deteriorated as well- mine is
still OK after 3 years of extensive use.

At 06:52 AM 6/20/2005, you wrote:



------------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I have been given the unenviable task of having to decide between a digital
camera system and a digital imaging plate system (the Ditabis system) to
enable our older TEMs to become fully digital imaging systems. We have Jeol
1200EX and 100CX instruments and a Phillips CM10. Has anyone had experience
of the Ditabis system? In the UK there is only one currently in use, and I
have contact with that person. Our samples are biological samples, some
routine clinical specimens, but also research samples and cell culture
samples. We currently use Kodak em film 4489 and require the replacement
system to have comparable resolution.
Any comments would be gratefully received.

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095

Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu






From MicroscopyL-request-at-ns.microscopy.com Fri Jun 24 10:31:46 2005



From: Microscopy Today :      microscopytoday-at-tampabay.rr.com
Date: Fri, 24 Jun 2005 11:31:00 -0400
Subject: [Microscopy] Microscopy Today July 2005 Table of Contents

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Listers,

Here is the July 2005 Microscopy Today table of contents. I will close
the subscription list for this issue on Thursday July 7, 2005.

Microscopists in North America and MSA members anywhere may have free
subscriptions. Anyone else may subscribe for US$35 per year (to
PARTIALLY cover postage). All subscriptions at
http://www.microscopy-today.com Thank you.

Ron Anderson, Editor
===============================
A Closer Look at the Nuclear Pore Complex
Stephen W. Carmichael, Mayo Clinic

An Innovative Method for Imaging and Chemical Analysis of Wet Samples in
Scanning Electron Microscopes
Irit Ruach-Nir, QuantomiX Ltd, Rehovot, Israel

An Aberration Corrected (S)TEM Microscope for Nanoresearch
S. Kujawa, B. Freitag, D. Hubert, FEI Company, Eindhoven, The Netherlands

Chemical Imaging with Infrared Microscopy
William J. McCarthy, Thermo Electron Corporation, Madison, WI

A Novel High-Speed Optical Scanning Microscope for Routine Clinical
Applications
D. Frekers1,2, I. Aksit1,2, V. Pilipenko1, K. Bünger;2 1Westfälische
Wilhelms-Universität Münster, 2MedXP-GmbH, Gelsenkirchen, Germany

Easy Guide to Calibrating TEM’s and STEM’s
John McCaffrey, Norrox Scientific Ltd. Ottawa, Canada

Embedding Small Specimens for TEM Examination
Paul Webster, House Ear Institute

The Total Release Method for FIB In-Situ TEM Sample Preparation
T.M. Moore, Omniprobe, Inc., Dallas, TX

EBSD Sample Preparation: Techniques, Tips, and Tricks
Matthew M. Nowell1, Ronald A. Witt2, and Brian W. True1
1EDAX-TSL, Draper, UT, 2EBSD Analytical, Lehi, UT

Funding Opportunities for Acquiring Equipment from Federal Granting
Agencies – Part III: IMR
Session Chair: Debby Sherman, Purdue University

M&M 2004 Core Facility Management – Part IV: Q & A
Moderator: Debbie Sherman

Industry News

NetNotes



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 24 12:46:38 2005



From: Irma Law :      rqmzaluzec-at-microscopy.com
Date: Fri, 24 Jun 2005 13:46:06 -0500
Subject: [Microscopy] Presenting Loans in 24 hours.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Howdy

Homeowner

You have been pre-approved for a $453,427 Home Loan at a 3.79 Fixed Rate.
This offer is being extended to you unconditionally and your credit is in no way a factor.

To take Advantage of this Limited Time opportunity

All we ask is that you visit our Website and complete
The 1 minute post Approval Form

http://9o1q70.mortgage-ebrokers.com/3/index/ryn/IgoFHT

Look Foward to Hearing From You,

Irma Law



From MicroscopyL-request-at-ns.microscopy.com Fri Jun 24 12:57:50 2005



From: claire haueter :      lukeclaire-at-yahoo.com
Date: Fri, 24 Jun 2005 10:57:04 -0700 (PDT)
Subject: [Microscopy] Repair of MT5000 Sorvall Ultramicrotome

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Listers,

Is there anyone out there in the Texas Medical Center
area or near vicinity that you know who can help
repair a MT5000 Sorvall ultramicrotome? It was
received as a gift but is obviously not working right
now. The person in charge of this thinks it might be
just a belt.

Anyone you can recommend?

T.I.A.,

Claire

Claire Haueter
EM Tech, Sr.
Integrated Microscopy Core
Baylor College of Medicine


__________________________________________________
Do You Yahoo!?
Tired of spam? Yahoo! Mail has the best spam protection around
http://mail.yahoo.com


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 24 13:15:41 2005



From: Steve Chapman :      protrain-at-emcourses.com
Date: Fri, 24 Jun 2005 19:18:44 +0100
Subject: [Microscopy] Judy Murphy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi

Could Judy Murphy please contact me?

Steve Chapman
Senior Consultant Protrain
For electron microscopy consultancy and training world wide
Tel +44 1280 816512 Fax +44 1280 814007
Mobile +44 7802 966067 Web www.emcourses.com


From MicroscopyL-request-at-ns.microscopy.com Fri Jun 24 17:36:28 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Fri, 24 Jun 2005 15:35:37 -0700
Subject: [Microscopy] Re: Re: Re: TEM digitisation: digital plates vs digital camera systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Jun 23, 2005, at 6:15 PM, Sergey Ryazantsev wrote:

} This is very common mistake thinking that more pixels gives you 'more'
} resolution. If your sample has, let say 10 nm details and nothing
} smaller, I would expect that quite decent 'resolution' on this
} particular sample would be if you utilize something like 3 pixels per
} 10 nm =} 3.3 nm/pixel. You simply don't need more pixels if there is
} nothing smaller on your sample. Adding more pixels you will basically
} add more noise to the image and that's it. Ultrathin sections of
} biological material, which I called 'biological samples' have quite
} poor resolution because of fixation, dehydratation plastic embedding
} and so on. Therefore, you will just spoil those 110M of pixels...
} It's like some people expected 3A resolution (resolution of the scope
} itself) on 80 nm thick poorly infiltrated chattered plastic sections
} of biological material, no way: 100A may be...
}
} Again, even on my 3.3GHz Pentium 5 with 2 Gb 3400 RAM computer, I do
} see deterioration in performance if I am working on couple 50Mb
} pictures, so you really need super computer to handle your 110 Mb
} pictures. And I am not talking about printing those images: on my
} networked color Xerox Phaser 7700GX with a lot of memory (maximum
} permitted) it takes about 20-30 sec to sent 10-20 Mb image (and more
} than minute to print at 1200 dpi). Your files will simply kill all our
} UCLA network! I think, you guys have better network, probably
} dedicated to handle such giant files.... So, I am just reasoning, why
} from my prospective, there is not much use of image plates in
} biological applications. I am glad that Ditabis is using solid state
} lasers - it definitely will add some juice to the system. By the way,
} how long it takes to scan one plate at maximum resolution?
}
} I do belive that image plate technology has a great potential in
} material science applications when huge dynamic range is critical.
} Have a great day, Sergey
}
Dear Sergey,
I agree that just having more pixels does not show smaller details
than are present in the specimen; however, if the mag is adjusted so
that the smallest details present are, as you suggest, about 3 pixels
across, then the more pixels you have, the larger the field of
view--which may be important. Another way of thinking about this is
that for larger pixels, as occurs with either image plates or CCDs, you
need a higher mag to see the same details, so you will have a smaller
field of view.
For a conventional fixed and stained biological plastic section, there
is no point whatever in having a magnification so large that you can
see details within the clumps of stain (typically ~2 nm), and there may
be no reason to have a mag high enough to distinguish individual stain
clumps (so even the smallest details in such a specimen may be of no
interest), but you still may want to see small gold clusters, e.g., on
antibodies or other labels, over a large piece of tissue, in which case
one would need high enough mag to identify the gold (or even to
distinguish among several sizes of gold) and a large enough field to
cover the entire region of interest in the specimen. It is in this
type of study that film is still ahead of IPs and CCDs due to the
higher resolution--especially with 4489.
Again, when you are talking about the file size that your computer can
handle (today; who knows about tomorrow), this may not be relevant to
the kinds of qualitative work for which film has an advantage. If one
wants to digitize the image, the advantages of film, even for the kind
of qualitative work I mentioned above, become much less pronounced,
since the scanner has larger pixels than the film, and it introduces
noise. In order to produce a 60,000 by 90,000 pixel image from a 6.5 x
9 cm film, the pixel size of the scanner would have to be 1um, and I
have never seen a 25,000 dpi scanner. In fact, it would not surprise
me at all if a CCD or IP with a 15 um pixel size gave a better quality
image than film scanned with a 10 um pixel size or even a 6 um (=4000
dpi, which is common) pixel size.
Finally, there is at least one biological application for which the
large dynamic range of IPs is useful: quantitative electron
diffraction.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Fri Jun 24 19:32:09 2005



From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Fri, 24 Jun 2005 17:32:07 -0700
Subject: [Microscopy] Re: Re: Re: Re: TEM digitisation: digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Bill
I completely agree with you: there are a number of application when large
view field is important and then old friend film or image plate may be a
solution. My point was that because of very low actual resolution on
ultrathin sections in particular, one don't need too much pixels to show
sample's details properly. For instance, my BioScan 600W camera being 1
Mpix has a view field size bigger than film - it was a whole point for
purposing that particular camera. So, I actually, have bigger than on the
film field size on my digital images. The price for that - low pixel
conts. But - I don't need much pixels to represent trustfully what I have
on my sections. It includes immuno-staining. I need to clarify: top mount
cameras have lover magnification than on the film, therefore they have
visible larger view area. In my particular case, my microscope do not
permits to take pictures at lower than x3K magnification, but I am able to
take pictures at x1.5K on digital camera, so I have more "field". Keeping
in mind your arguments in favor of really large view field I still keep my
dark-room running and 50 sheets of film is in the microscope. They are
sitting in the scope 3rd year: nobody wants to use film since digital
camera arrived. This is just practical observation: my users don't need
large field and high pixels count! What they need - to generate images
quickly and as many as they need, to be able to print them immediately,
transfer them through existing network (not 110 Mb) - digital camera suites
their need perfectly. For some rare work, I still have film
available. This is my practical solution. I think, purposing digital
camera was my best investment by the way.

As for scanning EM films - you absolutely right - film is
superior! Personally, I prefer old friend - film. Have a great
weekend, Sergey

At 03:35 PM 6/24/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_____________________________________

Sergey Ryazantsev Ph. D.
Electron Microscopy
UCLA School of Medicine
Department of Biological Chemistry
and UCLA Institute of Nanotechnology
10833 Le Conte Ave, Room 33-080
Los Angeles, CA 90095


Phone: (310) 825-1144 (office)
(310) 206-1029 (Lab)
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu





From MicroscopyL-request-at-ns.microscopy.com Sat Jun 25 09:57:57 2005



From: LDM Microscopy :      ldmm-at-risc4.numis.northwestern.edu.edu
Date: Sat, 25 Jun 2005 09:57:16 -0500 (CDT)
Subject: [Microscopy] Re: Postdoctoral Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Postdoctoral Position Available

A postdoctoral position is available at Northwestern to work
on a couple of projects, one involving low-loss energy loss
spectroscopy of surface plasmons (aloof spectroscopy) in
collaboration with a number of other faculty, the other a range
of electron microscopy on oxide materials related to fuel cells.
There will also be opportunities to interact with students/postdocs
working on several other projects including oxide surfaces, atomic-scale
tribology, heterogeneous catalysis and precession electron diffraction.

A strong general background in transmission electron microscopy and energy
loss spectroscopy is required, and some experience in the specific
research areas would be useful.

Applicants should apply by email only to L - marks -at- northwestern.edu
including the names and emails of three referees and a CV not more
than 3 pages in length including refereed publications and presentations.
Large attachments will invoke the delete button.

-----------------------------------------------
Laurence Marks
Department of Materials Science and Engineering
MSE Rm 2036 Cook Hall
2220 N Campus Drive
Northwestern University
Evanston, IL 60201, USA
Tel: (847) 491-3996 Fax: (847) 491-7820
email: L - marks -at- northwestern . edu
http://www.numis.northwestern.edu
-----------------------------------------------



From MicroscopyL-request-at-ns.microscopy.com Sun Jun 26 09:24:20 2005



From: david.griffiths-at-veths.no (by way of MicroscopyListserver)
Date: Sun, 26 Jun 2005 09:23:37 -0500
Subject: [Microscopy] viaWWW: Need LM sections from hpma blocks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (david.griffiths-at-veths.no) from http://www.microscopy.com/MLFormMail.html on Sunday, June 26, 2005 at 03:40:39
---------------------------------------------------------------------------

Email: david.griffiths-at-veths.no
Name: David Griffiths

Organization: Norwegian School of Veterinary Science

Title-Subject: [Microscopy] [Filtered] Need LM sections from hpma blocks

Question: We have a number of tissue blocks embedded in Technovit 8100 and other hpma's, but we aren't set up for sectioning yet. Is there anyone with with spare capacity who can cut sections of these blocks for us, for payment of course? If so could you please indicate your interest and we can sort out the details. The typical block would be about 10mm x 10mm, and we would like sections at 1.5 - 2.0 microns, mounted on a glass slide. About 10 - 15 sections per block at the moment.
Thanks,
Dave Griffiths
Section of Anatomy
Norwegian School of Veterinary Science

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Sun Jun 26 12:45:08 2005



From: Carol Heckman :      heckman-at-bgnet.bgsu.edu
Date: Sun, 26 Jun 2005 13:43:52 -0700
Subject: [Microscopy] Re: Re: Re: TEM digitisation: digital plates

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sergey-
Actually, there are two additional cases where the 3.3 nm/pixel
wouldn't help you with a problem. One is where you are trying for
automated extraction of the images of 5-nm gold particles, in order
to overlay them on the image. We have a procedure for this, and you
do have to have 3 pixels across the 5-nm image (minimum). It is
unpublished as yet (I am trying to get my former student to add it to
our on-line protocols). Another case where one would need } 3 pixels
across a feature is if the curvature of a contour is to be measured.
Here, pixellation artifact prevents any meaningful measurements from
being taken. To be specific, with a square array, one can only
measure angles at 90 degrees apart (straight to right and left, up
and down) and at 45 degrees apart (the corners). It is not possible
to derive a meaningful measure under these circumstances, and the
solution is to take more pixels and drop two between those subjected
to the final measurement. We published this solution in 1980 (if I
recollect correctly).

In any case, there are problems where lots of pixels in the array are
essential. The problem happen to be in the arena where the image
processing can be automated to effect, and that is the main advantage
of taking a digital image i.m.h.o. So the deck is not stacked
against the Ditabis technology -- the field just hasn't caught up
with the potential applications of the digital technology.
Carol
P.S. No commercial interest in Ditabis or its competitors.

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
__
Carol A. Heckman, Ph.D.
Professor of Biological Sciences
Director, Center for Microscopy & Microanalysis
Bowling Green State University, Bowling Green, OH 43403
fax: (419) 372-2024 email: heckman-at-bgnet.bgsu.edu
http://www.bgsu.edu/departments/biology/facilities/MnM
___________________________________________________________________________


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 08:13:15 2005



From: morten.laanre-at-imbv.uio.no (by way of MicroscopyListserver)
Date: Mon, 27 Jun 2005 08:12:33 -0500
Subject: [Microscopy] viaWWW: Zeiss 40x "antiflex" objective

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Below is the result of your feedback form (NJZFM-ultra-55). It was submitted by (morten.laanre-at-imbv.uio.no) from http://microscopy.com/MicroscopyListserver/MLFormMail.html on Monday, June 27, 2005 at 05:20:55
---------------------------------------------------------------------------

Email: morten.laanre-at-imbv.uio.no
Name: C.M.M. Laane

Organization: Institute of Molecular Bioscience,UiO Norway

Title-Subject: [Microscopy] [Filtered] MListserver:

Question: Does anyone have a Zeiss 40x "antiflex" objective for 160mm tube length for sale.Old catalog no. 46 17 54 (or 46 17 53) ?
The lens is needed for studies of cell attachment to substrate (surface).

Sincerely: C. Morten Motzfeldt Laane
Institute of Molecular Bioscience (IMBV)
University of Oslo, Norway
P.o.Box 1041, N-0316 Blindern, Norway

---------------------------------------------------------------------------


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 13:49:34 2005



From: Sergey :      sryazant-at-ucla.edu
Date: Mon, 27 Jun 2005 11:48:28 -0700
Subject: [Microscopy] Re: Re: Re: Re: TEM digitisation: digital

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Carol
I absolutely agree: there are lot of cases when you need more pixels. From
another hand, if you need more pixels for automatization - image plate
technology is not the best: you need manually load/unload plates, load them
into scanner, remove from scanner, load into cassette, load in microscope
etc... and after that you'll have 50 images without proper identification
(not everyone has last model microscopes). Keep in mind that scanning
itself at max resolution will take a while. I was asking Bill Miller a few
times how long it would take to scan a single plate. No answer. I would
imagine, it's comparable with scanning time on phosphorimager- about 20 min
3x4" area. If so (correct me Bill if I wrong) - 16 hours for 50
plates... Good digital camera is much, much better for automatization. I
am not even speaking about such wonderful features as "auto-tune", when
microscope automatically aligned at degree human eye could not. You also
could combine digital camera with energy filters etc - this is a way for
automatization. I used to work on new Hitachi microscope (don't remember
the model, 7200?) - they integrated 3D reconstruction from tilted series
into hardware - you just press knob and microscope (with digital camera)
made tilt series and reconstruct 3D image from it basically in real time -
you may not do anything like that with image plates, I am sorry Bill. Have
a great day. Sergey

At 01:43 PM 6/26/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 16:05:22 2005



From: Al Harmon :      aharmon-at-mvainc.com
Date: Mon, 27 Jun 2005 17:09:26 -0400
Subject: [Microscopy] ultra-microtomy of water soluable crystals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I have some very tiny spheres (3-5 microns in diameter) that I need to
obtain a thin section for TEM analysis, they are water soluable and
crumble easily. What I have done already is to mount them on blank stubs
with Extec and microtomy with limited results. Since they can't be
floated onto water I have been catching them dry onto carbon coated
copper TEM grids. They move quite a bit under the beam and the
crystalline interior crumbles.
Any thoughts would be appreciated.

Al Harmon
MVA Scientific Consultants, Inc.
"aharmon-at-mvainc.com"



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 16:12:37 2005



From: Bill Miller :      microbill-at-mohawk.net
Date: Mon, 27 Jun 2005 17:11:56 -0400
Subject: [Microscopy] Plate scanning time

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sergey - you must have missed the scanning time e-mail, I sent it to
you personally as you requested - we've had so many with the same
subject it may just gotten lost in the clutter... 2.5 minutes at 15
um and ~10minutes at 7.5um or about 1 hour to load, read and unload
the full load of 20 plates - less plates or lower resolution means
less time, of course..

A digital camera is certainly faster but if you need the resolution,
as she has indicated, they are quite costly - 4k x 4k, 14um pixel
cameras (only half the pixels of the 15um plate scanner) cost upwards
of $200k. I sell cameras for tomography as well and NOBODY does
tomography fast enough to have it qualify as "real time" . Even
the TVIPS 1k camera with } 12 frames per second takes quite some
time to do a full tilt series ( usually over 100 images). That said,
I'd never sell a plate system to anyone who was doing tilt series
topography - it's totally the wrong tool. Any application that
requires hundreds of images is best done with a CCD camera rather
than plats or film.


} -------------------------------------------------------------------------------
}
} Carol
} I absolutely agree: there are lot of cases when you need more
} pixels. From another hand, if you need more pixels for
} automatization - image plate technology is not the best: you need
} manually load/unload plates, load them into scanner, remove from
} scanner, load into cassette, load in microscope etc... and after
} that you'll have 50 images without proper identification (not
} everyone has last model microscopes). Keep in mind that scanning
} itself at max resolution will take a while. I was asking Bill
} Miller a few times how long it would take to scan a single
} plate. No answer. I would imagine, it's comparable with scanning
} time on phosphorimager- about 20 min 3x4" area. If so (correct me
} Bill if I wrong) - 16 hours for 50 plates... Good digital camera is
} much, much better for automatization. I am not even speaking about
} such wonderful features as "auto-tune", when microscope
} automatically aligned at degree human eye could not. You also could
} combine digital camera with energy filters etc - this is a way for
} automatization. I used to work on new Hitachi microscope (don't
} remember the model, 7200?) - they integrated 3D reconstruction from
} tilted series into hardware - you just press knob and microscope
} (with digital camera) made tilt series and reconstruct 3D image from
} it basically in real time - you may not do anything like that with
} image plates, I am sorry Bill. Have a great day. Sergey




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 16:35:31 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 27 Jun 2005 14:34:48 -0700
Subject: [Microscopy] Re: ultra-microtomy of water soluable crystals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Jun 27, 2005, at 2:09 PM, Al Harmon wrote:

} They move quite a bit under the beam and the crystalline interior
} crumbles.
}
Dear Al,
You could try carbon coating after the particles have been picked up
on the carbon-coated grids, and you could try a lower beam current.
The movement under the beam indicates either charging, heating, both,
or some other process--beam-induced specimen movement has not been
completely characterized--and increasing electrical and thermal
conductivity, or lowering the rate at which charge and/or heat
accumulates on the specimen will, to some extent, ameliorate the
problem.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 16:37:50 2005



From: Nestor J. Zaluzec :      zaluzec-at-aaem.amc.anl.gov
Date: Mon, 27 Jun 2005 16:37:06 -0500
Subject: [Microscopy] TEM Negative digitization times.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sergey

Stating that it takes 20 minutes to scan a negative at high
resolution without
specifics of what you mean or without comparisons to actual modern data
does not do justice to the purpose of this forum or the technological
advancements
that are in the scientific market today.

I have just gone to my older high resolution scanner (Leafscan 45)
here at ANL and
have digitized a 3 1/4 x 4 " negative at 1270 dpi ~ or a
resolution of 20 microns/pixel.
The image size was 5080 x 4128 pixels (21 Megapixels) and 16 bits deep.
The total time from the point that I mounted the negative in the holder
to when I had a digitized image on my screen was just under 3 minutes,
not 20 minutes. This is a fairly old scanner and I know that
newer ones are faster
and have higher resolutions (abeit not always with as good a bit depth).

One certainly could have gotten a similar image with a CCD camera by taking
multiple images (the number varying with the resolution of the camera)
but not in the same amount of time or cost. It would require
either a very large and expensive CCD camera or off-line stitching of
multiple images to merge and align the individual images together to
form an equivalent single image.

I use both technologies in my research, choosing the best technology
for the image
as appropriate, just as I would selecte the most appropriate
microscope to analyze
a specimen.

Let me remind everyone, that we all have the responsibility to make sure that
we are reasonably factual, accurate and up-to-date when we
point out the pros and cons of any technology. Using significantly out of
date or inaccurate information does not help disseminate useful knowledge
to the community, which is one of the purposes of this forum.


Nestor
Your Friendly Neighborhood SysOp

} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

--
===========================================
Dr. Nestor J. Zaluzec
Argonne National Lab
Materials Science Division/Bldg 212
9700 S. Cass Ave
Argonne, Illinois 60439 USA
Tel: 630-252-7901, Fax: 630-252-4798
Email: Zaluzec-at-aaem.amc.anl.gov
===========================================
TPMLab: http://tpm.amc.anl.gov
NEESLab: http://neestpm.mcs.anl.gov
MMSite: http://www.amc.anl.gov
===========================================

The box said ...
"This program requires Win 95/98/NT or better..."
So I bought a G3 Mac !

===========================================


From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 16:43:40 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Mon, 27 Jun 2005 14:42:57 -0700
Subject: [Microscopy] F30H objective aperture & low dose--solution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Lists,
Thanks everyone for the many helpful comments and suggestions. After
a series of investigations, we tried something that, in retrospect, was
obvious. When a smaller objective aperture was inserted, the problem
became less severe, leading to the conclusion that somehow the larger
objective aperture was out of position. Some of the possibilities are
that the aperture was not placed properly in the holder, that the foil
came loose from the body of the aperture, or that the tip was being
displaced when moved to the farthest-in position. In our case the we
still do not know exactly what occurred, but the apertures were
properly positioned and looked fine. This can occur in a microscope
where the aperture is in the back focal plane, as well as one like the
F30 where the aperture is not, so if you experience a similar problem
with your aperture not being centered in Focus state when it is
centered in Exposure state, see if this occurs just with one aperture
or is present for all the apertures.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 18:39:18 2005



From: Sergey :      sryazant-at-ucla.edu
Date: Mon, 27 Jun 2005 16:38:10 -0700
Subject: [Microscopy] Re: TEM Negative digitization times.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Nestor
In my message I was talking about image plates, not negatives. In my
speculation about time necessary to scan image plate I based on estimation
from scientific instrument, which is using very similar technology as a
Ditabis plate reader: phosphorimager. It uses basically the same image
plates and "imaging technology" is very similar if not identical. Latest
model of our Departmental phosphorimager Typhoon 9410 needs about 40 min to
scan 2x3" area with resolution of 10 um (see specification on instrument).
For 25 um, which perhaps comparable with Ditabis it would be 15-20
min. You may not compare optical scanner with image plate technology -
it's absolutely different instruments and principles! By the way, if you
will use good specialized optical scanner for difractogtramm (which gives
you optical density) - it will take much longer than 2-3 min to scan EM
size film, so I am quite sure that my approximation is quite right and this
is exactly why I was talking about that: many people would think similar to
you that "scanning is not a big deal". To get correct readings in OU is
basically a big deal even nowadays. "Household" scanners may not be used
for quantitative analysis - they are not calibrated and sometime not
linear. Their "color" sensor (depends from construction) may compromise
the data. In my message, I clearly stated that I was trying to obtain
technical data from Ditabis and they refuse to answer, so I have to
speculate instead using real data from manufacturer. Basically, even if it
takes according you just 3 min per plate - still, it will take 2.5 hours to
scan the standard set of 50 plates. Not very impressive to me. I am sorry
if I broke some rules AGAIN, I am naturally rule-breaker, I think most of
the science progress based on breaking somebody's rules (your boss
perhaps?). Have a good day, Sergey

At 02:37 PM 6/27/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 20:45:18 2005



From: Bill Miller :      microbill-at-mohawk.net
Date: Mon, 27 Jun 2005 21:44:37 -0400
Subject: [Microscopy] Re: Re: TEM Negative digitization times.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Sergey - just for comparison sake lets try to compare apples to
apples... figure out how long it would take to do 50 film images with
darkroom work - developing, drying , printing, drying - it's been so
long since I've had to actually do that I'd no longer feel confident
in even venturing a guess - but not minutes for sure

and then with your 1k camera shot to cover the same area as the film
at the same resolution as the plates with image tiling - that's
roughly 35-40 CCD images per 6k x 5k image plate (don't forget the
necessary overlap for the tiling) for a total of ~1800 CCD images --
you claimed that you took 100 images in 40 minutes so the math works
out to roughly 720 minutes (12 hours) BEFORE tiling

Two and a half hours to viewable images with the plates doesn't look
so bad from a time stand point when you look at the process
objectively - particularly since for most of the time the scanner is
doing all the work leaving you free to do something else. If you
don't need or want the resolution then it's rather hard to make a
rational comparison between low pixel count CCDs and imaging plates
since the image plates would not be the proper imaging device for
such an application anyway. No body is trying to talk anybody into
using the plates for anything other than what they are good for. If
I had to pull a 60 foot trailer I wouldn't pick a motorcycle to do it
with just because it was faster than a truck.

That said, the plates may still not be the best answer for many
application. When short turn around time is critical CCD cameras are
the answer. There is an endless list of applications where one
technology is clearly more suitable than another. Pick the one that
work for you.

Regards - Bill

At 07:38 PM 6/27/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Mon Jun 27 23:53:42 2005



From: Nestor J. Zaluzec :      zaluzec-at-aaem.amc.anl.gov
Date: Mon, 27 Jun 2005 23:52:58 -0500
Subject: [Microscopy] TEM Negative & Image Plate digitization times

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Bill

It is good to see factual information about scanning times for image plates. I also concur with your comments on using this technology for tomography.

The digitization times you presented are certainly consistent with my experiences using a high resolution high end negative scanners like the Leafscan 45, which is certainly not a run of the mill household scanner as the un-informed might presume.

As a followup, I would like to hear of comparisons using the older technology namely-negative drum scanners if anyone is still using this technology. I haven't used one of these used in years, and only vaguely remember using one over a decade ago here at ANL, it unfortunately has long since disappeared so I cannot test it out nor do I still have access to the original instrument specifications for comparison.

If anyone has additional real experience with plate or other high end negative scanners to add to this discussion it would be appropriate to post that data for the archives and to complete this discussion.

Finally, out of interest I checked on the specification of the Typhoon 9410 which was used as a basis for some scanning time estimates posted earlier in this discussion thread, which seemed unrealistic to me.

I was surprized to find out that the Typhoon 9410 is neither a negative nor image place scanner, but rather is a "high performance gel and blot imager ". Checking the technical specifications I see that while it has the same bit depth resolution as the Leafscan 45, it only has a uniformity of +/5% over the scan area, making this unsuitable for TEM work. The Typhoon manufacturer GE, in addition, makes absolutely no claims or has any application notes on its WWW site which indicate it is suitable for use with TEM negatives or image plates, nor is it apparently designed to be used for TEM application. It's scanning times are optimized for the detection of signal in gel/blot image data and not for data relevant to the discussion in this thread. In light of this fact, the performance characteristics estimated from an instrument like the Typhoon9410 and then extrapolated to TEM negative and/or image plate scanning are, in my opinion, a comparison which carries no technical merit.


Nestor
Your Friendly Neighborhood SysOp


Disclaimer: I have no financial interests in any of the imaging/scanning technologies discussed herein.



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 09:31:27 2005



From: Tom W Bargar :      tbargar-at-unmc.edu
Date: Tue, 28 Jun 2005 09:30:42 -0500
Subject: [Microscopy] EDS Detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html





A couple of questions. First, I would like to receive information from
vendors on their current models of elemental x-ray detectors. I would also
like to hear a comparison between the detectors that use liquid nitrogen
and those models that don't need liquid nitrogen cooling. Secondly, while
we hope to get a new SEM, we may not be able to, in that event I would like
to know if a EDX unit could still be fitted to our JEOL T220A SEM. As
always any information is appreciated from everyone.

Tom Bargar
Core Electron Microscopy Research Facility
University of Nebraska Medical Center
986395 Nebraska Medical Center
Omaha, NE 68198-6395
phone: 402-559-7347
tbargar-at-unmc.edu



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 11:25:01 2005



From: Andrew Johnstone :      johnstone-at-uky.edu
Date: Tue, 28 Jun 2005 12:17:12 -0400
Subject: [Microscopy] TEM - florescent bead labeling tissues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hello,

Checking if anyone knows of any new florescent bead labeling techniques when marking specific sites on tissues. Specifically, something that will not dissolve in PO. A polystyrene replacement?

Thanks

~~~~~~~~~~~~~~~~~~~~~~~
Andrew F.M. Johnstone
Deptartment of Biology
Cooper Lab
University of Kentucky
email: johnstone-at-uky.edu



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 13:25:49 2005



From: Bill Tivol :      tivol-at-caltech.edu
Date: Tue, 28 Jun 2005 11:24:03 -0700
Subject: [Microscopy] Re: TEM Negative & Image Plate digitization times

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


On Jun 27, 2005, at 9:52 PM, Nestor J. Zaluzec wrote:

} If anyone has additional real experience with plate or other high end
} negative scanners to add to this discussion it would be appropriate
} to post that data for the archives and to complete this discussion.
}
Dear Nestor,
I have had experience with scanners ranging from a CCD used to image
the light transmitted through a negative, through a linear diode array,
and a HiScan drum scanner to a Perkin Elmer microdensitometer. For
images, where generally the intensities do not vary rapidly over small
distances, and where the OD is centered around 1 and ranges from a few
tenths to, say, 2, any of the scanners will give a decent
representation of the image, and the fidelity of quantitation depends
on the number of pixels, the response time of the electronics, and the
stability of the illumination (among other variables). For electron
diffraction patterns, where the range of ODs is from ~0.001 to ~4, and
can vary greatly over distances on the order of tens of um, accurate
quantitation cannot be obtained unless the illuminated area of the
negative is the same as the scanning pixel. This is due to the fact
that only a small fraction of the light transmitted through adjacent
areas and scattered in the detector housing will give a falsely low OD
reading. For quantitation of ED patterns, only a microdensitometer,
such as the Perkin Elmer, will give sufficiently accurate data, and
then only at a fairly slow scan speed and small pixel size. I am not
up to date on the latest improvements in scanning times, but back in
the day such a scan would take hours for each negative. I have no
relationship to Perkin Elmer except as a former satisfied user.
Yours,
Bill Tivol, PhD
EM Scientist and Manager
Cryo-Electron Microscopy Facility
Broad Center, Mail Code 114-96
California Institute of Technology
Pasadena CA 91125
(626) 395-8833
tivol-at-caltech.edu




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 13:52:43 2005



From: Sergey :      sryazant-at-ucla.edu
Date: Tue, 28 Jun 2005 11:51:50 -0700
Subject: [Microscopy] Re: TEM Negative & Image Plate digitization times

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Nestor
You need to read my postings: I did not claim that Typhoon is a "scanner"
for negatives or have something to do with EM. Nevertheless, it's the best
on the market reader for "IMAGE PLATES". Image plate technology used in
many areas when relatively high energy particles (alpha/beta particles etc)
or gamma need to be detected. It's commonly used in molecular biology when
people used radioactive (32P, 14C, 3H) labelling, DNA for instance. You
separate your stuff on the gel and detect radioactivity using image plate.
In this area image plate technology is well established and
respected. Because (and I mentioned it in my original posting) image plate
technology is the same for EM and for other application it's legit and
even interesting to compare performance of different instruments- "image
plate readers" (correct name for it). From another hand I could not
understand why you are trying to compare optical scanners with image plate
readers? I think it's simply not legit to make any bridge between (even
non-household) optical scanner and image plate reader because
of differences in principles of operation - you have to know better, you
read manual to Typhoon. Leafscan 45 based on my quick research on Internet
- it's very good professional scanner for general photography, it scans
color and B&W slides and transparencies etc - it's not specialized scanner
for EM, it's not a densitometer ether- similarly as Typhoon is not
specialized reader for EM. If you think you may compare the data from
non-EM general scanner and image plate readers, than, my comparison of
Typhoon and Diabis is even more legit! Typhoon will scan Diabis image
plates (as any other) perfectly, by the way. 5% uniformity - this is what
you may expect from image plate readers due image plate properties, I
guess, and this is another point to think twice before using this
technology. I have about 10 years experience with image plates and readers
(in molecular biology where this technology is proven to have numerous of
advantages), so I am quite confident that information I provided is correct
and it's not my fault if somebody do not read the original postings and
misinterpret it: may be it's just because my broken English? I am sorry
for that. Sergey

At 09:52 PM 6/27/2005, you wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 15:16:00 2005



From: Barbara Foster :      bfoster-at-mme1.com
Date: Tue, 28 Jun 2005 15:15:01 -0500
Subject: [Microscopy] Re: ultra-microtomy of water soluable crystals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Dear Al,

We've been working with NT-MDT and Leica on a new AFM/Ultramicrotome hybrid that could provide an answer to your dilemma. The fully automated system will be announced officially later this year but the manual version was shown at Cell Bio last December and will be at the SPM '05 in Sapporo this month. It allows you to cut then image in one system, without having to catch sections. The results can be fed into a 3D reconstruction program (NT-MDT has tried, successfully, with Media Cybernetic's 3D Constructor) to generate 3D images and resulting image analysis. The AFM uses local differences in elasticity and hardness (among other modalities) to create contrast. If you would like to send some samples, we can forward them to NT-MDT and have them run them for you. You can contact me off-line for details.

Hope this is helpful.

Best regards,
Barbara Foster

Microscopy/Microscopy Education
313 S Jupiter Rd, Suite 100
Allen, TX 75002
P: 972-954-8011
W: www.MicroscopyEducation.com

P. S.
Need a good general reference or light microscopy text? Call us today to learn more about "Optimizing LIght Microscopy". Copies still available through MME... even for class-room lots ... and we give quantity discounts. Call us at (972)954-8011.

At 04:09 PM 6/27/2005, Al Harmon wrote:


} ------------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 16:00:05 2005



From: Mardinly, John :      john.mardinly-at-intel.com
Date: Tue, 28 Jun 2005 13:59:16 -0700
Subject: [Microscopy] TEM Negative & Image Plate digitization times

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Another issue to consider in negative scanning vs. image plates, is that
only a fraction of the negatives would ever get scanned because
examination of the negatives with a magnifier (human scan?) typically
reveals that too many of the negatives are not perfect, due to focus,
drift, or some other glitch. I can't quote a number, but for 30 years it
always seemed too high, either due to my skill, or in later years to my
being too picky (or both). With image plates, every exposure needs to
get scanned to be observed.

John Mardinly
Intel

This opinion is the opinion of the author, not of Intel corporation.
-----Original Message-----
} From: Nestor J. Zaluzec [mailto:zaluzec-at-aaem.amc.anl.gov]
Sent: Monday, June 27, 2005 9:53 PM
To: microscopy-at-microscopy.com

Bill

It is good to see factual information about scanning times for image
plates. I also concur with your comments on using this technology for
tomography.

The digitization times you presented are certainly consistent with my
experiences using a high resolution high end negative scanners like the
Leafscan 45, which is certainly not a run of the mill household scanner
as the un-informed might presume.

As a followup, I would like to hear of comparisons using the older
technology namely-negative drum scanners if anyone is still using this
technology. I haven't used one of these used in years, and only
vaguely remember using one over a decade ago here at ANL, it
unfortunately has long since disappeared so I cannot test it out nor do
I still have access to the original instrument specifications for
comparison.

If anyone has additional real experience with plate or other high end
negative scanners to add to this discussion it would be appropriate to
post that data for the archives and to complete this discussion.

Finally, out of interest I checked on the specification of the Typhoon
9410 which was used as a basis for some scanning time estimates posted
earlier in this discussion thread, which seemed unrealistic to me.

I was surprized to find out that the Typhoon 9410 is neither a negative
nor image place scanner, but rather is a "high performance gel and
blot imager ". Checking the technical specifications I see that while
it has the same bit depth resolution as the Leafscan 45, it only has a
uniformity of +/5% over the scan area, making this unsuitable for TEM
work. The Typhoon manufacturer GE, in addition, makes absolutely no
claims or has any application notes on its WWW site which indicate it
is suitable for use with TEM negatives or image plates, nor is it
apparently designed to be used for TEM application. It's scanning
times are optimized for the detection of signal in gel/blot image data
and not for data relevant to the discussion in this thread. In
light of this fact, the performance characteristics estimated from an
instrument like the Typhoon9410 and then extrapolated to TEM negative
and/or image plate scanning are, in my opinion, a comparison whic!
h carries no technical merit.


Nestor
Your Friendly Neighborhood SysOp


Disclaimer: I have no financial interests in any of the imaging/scanning
technologies discussed herein.





From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 16:24:23 2005



From: John J. Bozzola :      bozzola-at-siu.edu
Date: Tue, 28 Jun 2005 16:23:43 -0500
Subject: [Microscopy] TEM: digital and/or film scope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Theoretical Question: if you were purchasing a TEM equipped with a
high quality (2k x 2k) digital camera would you still opt for the
film camera as well? Just wondering.
--
##############################################################
John J. Bozzola, Ph.D., Director
I.M.A.G.E. (Integrated Microscopy & Graphics Expertise)
750 Communications Drive - MC 4402
Southern Illinois University
Carbondale, IL 62901 U.S.A.
Phone: 618-453-3730
Email: bozzola-at-siu.edu
##############################################################


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 16:32:42 2005



From: jmkrupp-at-cats.ucsc.edu (Jon Krupp)
Date: Tue, 28 Jun 2005 14:23:19 -0700
Subject: [Microscopy] Congo red & protein fibrils?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

I am told by someone who visits my lab that protein fibrils treated with
congo red will show an apple green birefringence in crossed polars.

She has brought samples for us to look at, but we cannot see the green
birefringence. She 'knows' there are fibrils present by another type of
assay.

Has anybody heard of this technique? I think I can see the stained fibrils,
but nothing shows with the crossed polars. Maybe I don't know what I should
be looking for, maybe it is very subtle. Maybe it doesn't really work.

If you have heard of this technique and maybe know a trick or two, let me
know so I can help this person.

Thanks

Jonathan Krupp
Microscopy & Imaging Lab
University of California
Santa Cruz, CA 95064
(831) 459-2477
jmkrupp-at-cats.ucsc.edu




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 16:41:51 2005



From: Philip Oshel :      peoshel-at-wisc.edu
Date: Tue, 28 Jun 2005 16:41:07 -0500
Subject: [Microscopy] Microscopy texts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Microphiles,

I'm looking for a couple of EM texts, and having no luck with the
on-line used book sources:
M. A. Hayat, "Fixation for Electron Microscopy"
Reimer and Hawkes "Scanning Electron Microscopy : Physics of Image
Formation and Microanalysis"
Anyone happen to have an extra copy in good or better condition that
they might be willing to part with?

Phil
--
Philip Oshel
Supervisor, BBPIC microscopy facility
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)
http://www.ansci.wisc.edu/microscopy.htm


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 17:11:09 2005



From: Damian Neuberger :      neuberger1234-at-comcast.net
Date: Tue, 28 Jun 2005 17:10:25 -0500
Subject: [Microscopy] Microscopy texts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Phil,
If you want, for the price of mailing them, Hayat Principles and Techniques
of Electron Microscopy Biological Applications Vols 1 and 3, 1970 and 1973
respectively. Kind of old but unless someone wants them, they go into the
trash along with a bunch more.
Damian

-----Original Message-----
} From: Philip Oshel [mailto:peoshel-at-wisc.edu]
Sent: Tuesday, June 28, 2005 4:41 PM
To: Microscopy-at-microscopy.com

Microphiles,

I'm looking for a couple of EM texts, and having no luck with the
on-line used book sources:
M. A. Hayat, "Fixation for Electron Microscopy"
Reimer and Hawkes "Scanning Electron Microscopy : Physics of Image
Formation and Microanalysis"
Anyone happen to have an extra copy in good or better condition that
they might be willing to part with?

Phil
--
Philip Oshel
Supervisor, BBPIC microscopy facility
Department of Animal Sciences
University of Wisconsin
1675 Observatory Drive
Madison, WI 53706
voice: (608) 263-4162
fax: (608) 262-5157 (dept. fax)
http://www.ansci.wisc.edu/microscopy.htm





From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 17:16:24 2005



From: Tobias Baskin :      baskin-at-bio.umass.edu
Date: Tue, 28 Jun 2005 18:15:35 -0400
Subject: [Microscopy] Re: Congo red & protein fibrils?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Greetings,
The trick here is that you have to think in terms of
dichroism rather than birefringence. What happens with congo red (at
least in theory) is that it binds to the substrate in an oriented
way. So if you have a substrate that is oriented then the congo red
molecules themselves become oriented. Now congo red happens to be
good at absorbing linearly polarized light oriented along one axis of
the molecule and bad along the perpendicular axis. So a congo red
stained sample should act like a polarizing filter, itself. So to see
your apple green signal, what you do is to remove the analyzer from
the system, have just the polarizer (you can do it the other way if
you want, it doesn't matter). Then rotate the sample through at
least 90 degrees. You are using the sample to make "crossed
polarizers". The apple green should show up when the sample fibers
are at some defined orientation to the polarizer and the signal
should disappear at +/- 90 degrees to that orientation (no getting
more specific without knowing which way the fibers go and how the
congo binds to them).

You get green presumably becaue the congo red absorbs red and
blue better and so more green leaks through. Actually, polarizing
filters themselves are often made by aligning really strong crystals
that have dichroic abosorption, and they too have a green cast
because it is difficult to have no wavelength dependence to the
dichroism.

Keep in mind that congo red stained fibers are going to be a
whole heck of a lot less good at being a polarizing filter than a
real polarizing filter, so the green signal may be pretty faint. But
its tell tale appearance and disappearence at +/- 90 gives you a
story. You can play with using green light (or other wave bands) but
I think one's eyes are most senstivite to seeing differences in color
rather than intensity.

Hope this helps,
Tobias
}
}
} I am told by someone who visits my lab that protein fibrils treated with
} congo red will show an apple green birefringence in crossed polars.
}
} She has brought samples for us to look at, but we cannot see the green
} birefringence. She 'knows' there are fibrils present by another type of
} assay.
}
} Has anybody heard of this technique? I think I can see the stained fibrils,
} but nothing shows with the crossed polars. Maybe I don't know what I should
} be looking for, maybe it is very subtle. Maybe it doesn't really work.
}
} If you have heard of this technique and maybe know a trick or two, let me
} know so I can help this person.
}
} Thanks
}
} Jonathan Krupp
} Microscopy & Imaging Lab
} University of California
} Santa Cruz, CA 95064
} (831) 459-2477
} jmkrupp-at-cats.ucsc.edu


--
_ ____ __ ____
/ \ / / \ / \ \ Tobias I. Baskin
/ / / / \ \ \ Biology Department
/_ / __ /__ \ \ \__ 611 N. Pleasant St.
/ / / \ \ \ University of Massachusetts
/ / / \ \ \ Amherst, MA, 01003
/ / ___ / \ \__/ \ ____
http://www.bio.umass.edu/biology/baskin/
Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243


From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 18:05:14 2005



From: Aleksandr.Mironov :      Aleksandr.Mironov-at-manchester.ac.uk
Date: Wed, 29 Jun 2005 00:04:08 +0100
Subject: [Microscopy] Re: Congo red & protein fibrils?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

Hi,

To my knowledge green birefringence after Congo Red treatment appears
when your protein has a lot of b-sheets in its structure. Amyloids
formed from different proteins (APP, prion etc.) show such phenomenon.
So, perhaps, the sample you was looking at did not have amyloid
conformation.

Regards,
Aleksandr Mironov
EM Unit
University of Manchester


On 28 Jun 2005, at 22:23, Jon Krupp wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} I am told by someone who visits my lab that protein fibrils treated
} with
} congo red will show an apple green birefringence in crossed polars.
}
} She has brought samples for us to look at, but we cannot see the green
} birefringence. She 'knows' there are fibrils present by another type of
} assay.
}
} Has anybody heard of this technique? I think I can see the stained
} fibrils,
} but nothing shows with the crossed polars. Maybe I don't know what I
} should
} be looking for, maybe it is very subtle. Maybe it doesn't really work.
}
} If you have heard of this technique and maybe know a trick or two, let
} me
} know so I can help this person.
}
} Thanks
}
} Jonathan Krupp
} Microscopy & Imaging Lab
} University of California
} Santa Cruz, CA 95064
} (831) 459-2477
} jmkrupp-at-cats.ucsc.edu
}
}
}



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 18:50:11 2005



From: Sally Stowe :      Sally.Stowe-at-anu.edu.au
Date: Wed, 29 Jun 2005 09:49:10 +1000
Subject: [Microscopy] ] Re: TEM Negative & Image Plate digitization times-contrast range

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html


An evaluation of a Ditabis scanner a few years ago left the impression that
it was slightly less convenient to use than film, although now that less
film is being processed that may be changing. But the very great advantage
in our applications was the dynamic range rather than spatial resolution. 32
bit is something else.
Diffraction patterns easily showed features that would have had to be
carefully sought out using film. For biological work, IP plates offer the
possibility of much more frequent use of unstained material - looking at the
real tissue rather than a heavy metal incrustation of osmium, lead or
uranium, in embedded as well as frozen material. I would be interested to
know if anyone has been doing much along these lines?

IP is complementary to CCD cameras at the moment, rather than a practical
replacement. The immediacy of the CCD camera is such an advantage that its
hard to remember how we worked without it. My take on the "2K x 2K CCD vs
film camera" question is "2k by 2K...keep the film just in case. But once
4K by 4k or more comes in affordably..... time for some serious re-design of
the TEM!

My two cents worth..

Cheers
Sally



Dr SJ Stowe
Facility Coordinator
ANU Electron Microscopy Unit
ANU CRICOS#00120C



} -----Original Message-----
} From: Bill Tivol [mailto:tivol-at-caltech.edu]
} Sent: Wednesday, 29 June 2005 4:24 AM
} To: microscopy-at-msa.microscopy.com
} Subject: [Microscopy] Re: TEM Negative & Image Plate digitization times
}
}
}
}
} ---------------------------------------------------------------
} ---------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver

}
}
}




From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 19:05:37 2005



From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Wed, 29 Jun 2005 12:05:57 +1200
Subject: [Microscopy] Re: EDS Detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

}
} A couple of questions. First, I would like to receive information
} from vendors on their current models of elemental x-ray detectors. I
} would also like to hear a comparison between the detectors that use
} liquid nitrogen and those models that don't need liquid nitrogen
} cooling. Secondly, while we hope to get a new SEM, we may not be
} able to, in that event I would like to know if a EDX unit could still
} be fitted to our JEOL T220A SEM. As always any information is
} appreciated from everyone.
}


Whatever you do, think carefully about the detailed mechanical design of
whatever detector you end up buying, and, if possible, the manufacturer's
reputation for standing behind their products.

I have an approximately 5-year-old PGT Prism 2000, the sort which has an
adjustable dewar angle to allow it to be used at a variety of takeoff angles.

At 2 1/2 years old, its LN2 consumption had steadily increased to the point
where it had to be refilled every second day, so I returned it to PGT to have
the leak fixed and to be re-evacuated, for over a thousand bucks.

When I (eventually) got it back, the LN2 consumption was OK, but over the
next two years it steadily deteriorated again, quite spontaneously (no
warmups or other mishaps), and, to my extreme disappointment, PGT's
non-negotiable attitude is "it was OK when it left here, it must have
developed another leak".

This defies common sense --- we're talking here about a detector which
twice experienced a dewar vacuum deterioration, and whose leak was
apparently not fixed when it was returned to the factory.

I don't think it's a leaky window, as it's been on my SEM under vacuum 24/7
all the time, so my suspicion falls on the O-ring-sealed elbow.

Unfortunately, the PGT vacuum port is non-standard and is on the top
surface right underneath the dewar, and, even more ridiculously, the PGT
adaptor (US$500!!!) won't fit onto it if the dewar is tilted to be vertical for a
40 degree takeoff angle, so, you guessed it, AFTER evacuating the
detector, the bolts holding the elbow joint have to be slackened so that the
elbow can be adjusted to the correct angle! This, of course, placed the
bright shiny new vacuum at risk.

The guys in the excellent workshop to which I have access have now made
me an adaptor which will fit onto the untilted detector, and when I work up
the courage I will repump it myself.

Of course, everyone makes their own purchase decisions, but I will never
again buy a detector which:

- has an adjustable elbow
- has a non-standard vacuum port
- is made by a manufacturer who won't guarantee the vacuum integrity of
the dewar for five years following manufacture or factory repair (barring
accidents)
- is made by PGT.

In the meantime I fill the damned thing every day, including Saturdays and
Sundays, with considerable resentment.

cheers

rtch



--
Ritchie Sims Ph D Phone : 64 9 3737599 ext
87713
Microanalyst Fax : 64 9 3737435
Department of Geology email :
r.sims-at-auckland.ac.nz
The University of Auckland
Private Bag 92019
Auckland
New Zealand



From MicroscopyL-request-at-ns.microscopy.com Tue Jun 28 19:06:17 2005



From: David Elliott :      Elliott-at-arizona.edu
Date: Tue, 28 Jun 2005 17:05:32 -0700
Subject: [Microscopy] Re: TEM: digital and/or film scope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

YES
2x2 is good, but film is still better for some things. I just wrote a
proposal for a 'scope with both.
David


On Jun 28, 2005, at 2:23 PM, John J. Bozzola wrote:

}
}
} -----------------------------------------------------------------------
} -------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
} To Subscribe/Unsubscribe --
} http://www.microscopy.com/MicroscopyListserver
} On-Line Help http://www.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------
} --------
}
} Theoretical Question: if you were purchasing a TEM equipped with a
} high quality (2k x 2k) digital camera would you still opt for the film
} camera as well? Just wondering.
} --
} ##############################################################
} John J. Bozzola, Ph.D., Director
} I.M.A.G.E. (Integrated Microscopy & Graphics Expertise)
} 750 Communicati