From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Fri, 1 Oct 1993 15:18:05 -0500 (CDT)
Subject: Open for Business

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Microscopy Listserver/Mailreflector is
now running you may begin posting messages

Cheers- Nestor




From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Sat, 2 Oct 1993 11:03:48 -0500 (CDT)
Subject: TEM Gun Brightness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since I decided to bit the bullet and try to get this
system working let me post the first real message to
the Microscopy Listserver.

===============================================================

We have been experimentally measuring electron gun brightness
of LaB6 and CeB6 sources in our TEM's here and find that
values are factors of 4-5 times lower than TEM manufacturers claims.

For example, at 400 kV experimentally we determine values of
~ 2-3 x 10**6 A/cm**2/sr while values quoted are more typically
} 1 x 10**7 A/cm**2/sr. Small variances are found when one
adjusts filament temp, bias, wehneldt gap & filament but not
enough to make up the difference.

The brightness (b) is determined by experimental measurements in the
electron microscope using the following equation:

b = 4I/(pi**2d**2a**2)

where I = probe current pi=3.14159, d = beam diameter (FWTM), a =
incident beam divergence half angle. All parameters are quantitatively
measured in the instrument.

We have found no reports of similiar measurements in the literature,
except for the initial studies of source brightness done years ago
which of course everone quotes. None of which were done in TEM's,
but most were rather "bench" measurements. Has anyone else actually measured
the performance of their guns in a TEM (or SEM)?


Nestor J. Zaluzec
ANL EM Center




From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Sun, 3 Oct 1993 11:13:22 -0600
Subject: re: TEM Gun Brightness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Since I decided to bit the bullet and try to get this
} system working let me post the first real message to
} the Microscopy Listserver.
}
} ===============================================================
}
} We have been experimentally measuring electron gun brightness
} of LaB6 and CeB6 sources in our TEM's here and find that
} values are factors of 4-5 times lower than TEM manufacturers claims.

** specific stuff deleted **

I don't have any hard data on operating conditions, but we have been
wondering why LaB6 cathodes don't seem significantly brighter than tungsten
in our TEM. We had expected LaB6 to be like looking at the sun, but that's
not the case. Could it be that the setup needs to be tweaked more
critically?

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: Paul Goodwin :      pgood-at-inson.fhcrc.org
Date: Mon, 4 Oct 1993 08:30:20 -0700 (PDT)
Subject: Re: Dichroic mirrors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Mon, 4 Oct 1993, Franz Keller wrote:

} Hello everybody,
}
} I am looking for dichroic mirrors for epifluorescence, which are able to
} reflect several ranges of wavelengths and allow transmission of light in
} the ranges outside. Does anybody know of companies which are designing such
} mirrors? It would be a great help to me if I know whom to contact.
} Thanks a lot
} Franz Keller
} *--------------------------------------------------------------------*
} | Dr.Franz Keller Tel.:089-8578-3688 |
} | MPI f. Psychiatry Fax 089-8578-3939 |
} | Dept. Neuromorphology |
} | Am Klopferspitz 18a EMail |
} | W-8033 Martinsried keller-at-vms.biochem.mpg.de |
} *--------------------------------------------------------------------*


The most respected source right now is a company called Chroma
Technologies Corp. They are a bunch of techies that left Omega Optical to
pursue the realm of multiple pass dichroic mirrors for biomedical
research. They can be reached at:

Chroma Technologies Corp.
(802)257-1800 (voice)
(802)257-9400 (fax)

I've been told that Paul Nilman is a good person to talk to there.


Paul Goodwin








From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Mon, 4 Oct 1993 11:43:07 -0500 (CDT)
Subject: Gun Brightness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



John C. replies

} I don't have any hard data on operating conditions, but we have been
} wondering why LaB6 cathodes don't seem significantly brighter than tungsten
} in our TEM. We had expected LaB6 to be like looking at the sun, but that's
} not the case. Could it be that the setup needs to be tweaked more
} critically?

Minor clarification. The LaB6 values we measured are low relative
to calculations and claims however, they are better than
W values! I'll dig up the relative comparisons and post
them later.

Are you running your LaB6 at 1st or 2nd x-ver of the source?

Nestor Zaluzec
ANL EM Center






From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Mon, 4 Oct 1993 11:19:24 -0600
Subject: Re: Gun Brightness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} John C. replies
}
} } I don't have any hard data on operating conditions, but we have been
} } wondering why LaB6 cathodes don't seem significantly brighter than tungsten
} } in our TEM. We had expected LaB6 to be like looking at the sun, but that's
} } not the case. Could it be that the setup needs to be tweaked more
} } critically?
}
} Minor clarification. The LaB6 values we measured are low relative
} to calculations and claims however, they are better than
} W values! I'll dig up the relative comparisons and post
} them later.
}
} Are you running your LaB6 at 1st or 2nd x-ver of the source?
}
} Nestor Zaluzec
} ANL EM Center

At the moment, we're not running LaB6. But the reasons why are for another
post. :-) I have always run LaB6 at the first maximum brightness I get
to, (1st X-over?) when illumination from the sides of the crystal merge
with that from the tip.

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: CSENCSITS-at-ANLEMC.MSD.ANL.GOV (ROSEANN CSENCSITS (708) 252-4977, -7902)
Date: Mon, 4 Oct 1993 13:50:48 -0500 (CDT)
Subject: LaB6-TEM-1st and 2nd cross over

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone with a LaB6 in a JEOL TEM ever see a second cross over?
I see them in Phillips microscopes but not in JEOL microscopes.
What is your experience?

Roseann Csencsits
ANL 708-252-4977
or 708-252-7902




From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Mon, 4 Oct 1993 13:29:03 -0600
Subject: Re: Gun Brightness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Regarding first and second x-ver of the LaB6--Do you have a JEOL or
} Phillips TEM? I have never seen the second cross over in a JEOL TEM.
}
} Roseann Csencsits
} Argonne National Lab
} 708-252-7902

I used a Philips 400 with LaB6 years ago when I was a graduate student. I
don't remember seeing a second brightness peak. When I got to CSU last
summer, our JEOL 2000 had LaB6, but we couldn't check the vacuum with the
original set of guages. We had to add a Penning guage that would give an
accurate readout to the 10\-7 range. Before we went back to tungsten, I
don't remember seeing that second peak.

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Tue, 5 Oct 1993 14:56:17 -0500 (CDT)
Subject: Apologies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Everyone


Seems like we have a fair number of people who are not
reading the subscription "destructions". In it I ask
new subscribers to reply to the listserver-at-anlemc.msd.anl.gov
to verify their address before I add them to the list.
Most of you did this correctly. :-)

I did this to minimize the possiblity of "bad addresses"
causing Email bounce where everyone on the list would
receive copies of "undeliverable" mail messages. for days
on end.

However a number new subscribers appear to be replying to
Microscopy-at-anlemc.msd.anl.gov. I'll try to make the
message clearer or sort out another way to check out/confirm
the addresses.

Please bear with it for a few more days.

Thanks - Nestor

P.S. I've have gotten 2 bad Email bounces both from
Germany, where the message repeated ~ 10 times/day for a
few days, so the extra work did stop that problem, only to
create the other. :-(






From: anandamu-at-scf.usc.edu (Anand)
Date: Tue, 5 Oct 1993 14:57:01 -0700 (PDT)
Subject: Seeking Postdoc/Industrial R&D Microscopist position.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

University of Southern California
Los Angeles CA 90089 Ph: (213) 740-1992
FAX: (213) 740-7797
X-Mailer: ELM [version 2.4 PL21]
Mime-Version: 1.0
Content-Type: text/plain; charset=US-ASCII
Content-Transfer-Encoding: 7bit
Content-Length: 1352

I am seeking a challenging position as a research Electron
Microscopist, so I thought this group would be appropriate to post
this article.

I graduated from the University of Southern California in May 1993
with a Ph.D. in Materials Science. As a graduate research associate, I
was involved extensively with conventional and high-resolution
transmission electron microscopy in research projects relating to
deformation twinning, martensitic transformations, and studies of
ordered and premartensitic phases. I have considerable experience in
working with shape memory alloys, intermetallic compounds, titanium
alloys, superalloys and ceramics. Further, I also have considerable
research experience with SEM, EDS, x-ray diffraction, AES and
metallography. With my present experience, I am confident I will be
able to undertake projects relating to research and development of a
wide variety of materials. I would certainly appreciate an opportunity
to discuss with you how I could contribute to you/your organization.
Please do not hesitate to call, or send email to me.

Sincerely,

ANANDA S. MURTHY

Center for Electron Microscopy EMail: ANANDAMU-at-USC.EDU
& Microanalysis Phone: (213) 740-1992
University of Southern California Fax: (213) 740-7797
Los Angeles, CA 90089-0101






From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Wed, 6 Oct 1993 9:19:27 -0500 (CDT)
Subject: Kikuchi Patterns

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Good programs have several different definitions:
do you mean commerical or free/shareware.

There are a few programs in the EMMPDL library under
the CBED directory which you can access. For a listing
send an Email message to

EMMPDL-at-ANLEMC.MSD.ANL.GOV

in the Email message enclose the command

Send HELP

you will receive information on how to use EMail to
access abstracts, directories and complete source code files.



Nestor Zaluzec
ANL EM Center




From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Wed, 6 Oct 1993 9:38:09 -0500 (CDT)
Subject: Resume's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



All Microscopy Users

This mailing list was not intended as a place to
post resume's to distribute to the world. Please
refrain from such usage. If you will recall in
the Welcome message it was pointed out that this
was to be a discussion/question/answer/information
forum.

If you wish you may post this type of information (ie. resume's)
instead on the MSA BBS system (accessible by Internet & modem)
which (if you are a member of MSA) has a positon/job
placement service available to members at no charge.

Posting of Position/Job openings is stretching the
limits of what the intention of this list is
supposed to address, however, I am willing to consider
this as potentially within the bounds. What is the
opinion of the readers on this last issue (i.e.
positon/job openings in the Microscopy community)


Nestor J. Zaluzec
ANL EM Center






From: COOK-at-ANLEMC.MSD.ANL.GOV
Date: Wed, 6 Oct 1993 12:28:14 -0500 (CDT)
Subject: RE: Resume's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310061627.AA00397-at-pascal.mrd.bldrdoc.gov}

"Microscopy" could quickly fill with resumes and job openings. I vote that
these notices should not be allowed in this forum.




From: tim-at-phlogiston.nist.gov (Tim Foecke)
Date: Wed, 06 Oct 1993 13:06:32 -0400
Subject: Re: Resumes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



A vote against resumes. But a vote for open positions.

Tim Foecke
tfoecke-at-nist.gov





From: Murray Foster :      vital!unixmailgate!murray-at-uunet.UU.NET
Date: 6 Oct 93 15:41:47 U
Subject: Re: Resume's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reply to: RE} Resume's
No resumes please.






From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Wed, 6 Oct 1993 16:18:22 -0500 (CDT)
Subject: position openings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: modb%BANRUC60.BITNET-at-ANLVM.CTD.ANL.GOV
Date: Wed, 6 Oct 1993 17:27:24 +0100
Subject: Re: Resume's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm strongly in favour of this "positon/job opening" ad's.
For us in Europe, it is very hard to get that kind of information
in time.

Marc Op de Beeck.
University of Antwerp (RUCA)
BELGIUM
modb-at-banruc60.bitnet





From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Wed, 6 Oct 1993 16:16:34 -0500 (CDT)
Subject: no resume's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: Gerry Little :      ANGJL-at-medicine.newcastle.edu.au
Date: Thu, 7 Oct 1993 08:14:44 GMT+1000
Subject: Re: welcome to microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


G'day,
Yes your message got through and my address is correct.
Regards,
Gerald Little.
Dr Gerry Little | Ph (049) 215618
Discipline of Anatomy | Fax (049) 216903
Faculty of Medicine | Email ANGJL-at-medicine.newcastle.edu.au
University of Newcastle




From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Wed, 6 Oct 1993 17:45:15 -0600
Subject: Re: Resume's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310062328.AA14275-at-enet-gw.pa.dec.com}

} Posting of Position/Job openings is stretching the
} limits of what the intention of this list is
} supposed to address, however, I am willing to consider
} this as potentially within the bounds. What is the
} opinion of the readers on this last issue (i.e.
} positon/job openings in the Microscopy community)
}
}
} Nestor J. Zaluzec
} ANL EM Center

Just for saving bandwidth and not inconveniencing uninterested people, I'd
say posting a message that states you are looking for a position, or have a
position open, is fine. This can be done in a few short sentences. Make
the details available privately.

I think distribution of an announcement to this very specialized group of
readers is completely appropriate, and even an important function of this
group.

Just my opinion.

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: ZALUZEC-at-ANLEMC.MSD.ANL.GOV (Nestor J. Zaluzec (708)-252-5075, -4964)
Date: Wed, 6 Oct 1993 21:06:36 -0500 (CDT)
Subject: resume opinion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: tim-at-phlogiston.nist.gov (Tim Foecke)
Date: Thu, 07 Oct 1993 10:25:22 -0400
Subject: Re: STM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We are in the same process. The machine you buy is very tightly related to
the type of work you need to do. Do you need vacuum? Are you going to be
using STM alone, or also AFM, MFM, LFM, M.O.U.S.E., ....? Do you want to
crack it open and make your own code? The Nanoscope III seems to be the better
for versatility and routine use, while Topometrix gives you the source code
of the software so you can do modifications.

Need more specifics.

Tim Foecke, NIST





From: Paul Goodwin :      pgood-at-inson.fhcrc.org
Date: Thu, 7 Oct 1993 15:41:30 -0700 (PDT)
Subject: Re: Re resumes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310071557.AA18638-at-enet-gw.pa.dec.com}



On Thu, 7 Oct 1993, John Crum wrote:

} To all:
}
} I counted no less than fifteen mail messages on the subject of resumes, in
} the last day. I think this is far more messages than the resumes you'll see
} posted in the next six months.
}
} We all have control over what we read. If you don't want to read another
} resume, just delete the message. This, of course, assumes that the
} responsible resume poster will use the subject line and state that the
} message is a resume.
}
} I vote that this is a valid subject (and a much needed service) on this
} list. With one ground rule, and that is that the writer must state in the
} subject line that the message is a resume.
}
} Let's face it, looking for a job is tough, especially in our tough econonmic
} times.
}
} John Crum
} San Diego State University
}

Ditto-

Paul Goodwin






From: Zaluzec-at-anlemc.msd.anl.gov (Nestor J. Zaluzec)
Date: Thu, 7 Oct 1993 15:41:30 -0700 (PDT)
Subject: Updates and Messages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: microscopy-at-anlemc.msd.anl.gov

G'day Microscopy Subscriber's

The system has been down for most of Thurs. while
I did network upgrades, hope this hasn't caused too
much trouble.

I've corrected the welcome notice about replies to the list
and tried to make the directions VERY CLEAR
so most of you should not be receiving copies
of confirmation notices any longer.

For those that have asked we currently have ~ 250 subscribers,
not all of whom have reconfirmed yet.

Secondly, when replying to a real comment/question
posted on the list please make sure that the
message you send is addressed to:

MICROSCOPY-at-ANLEMC.MSD.ANL.GOV

and not just a reply to the sender. There are many
different type of Email systems out there and not all
of them sort out the To;, From: CC: of the Email header.
in the same way. Thus you may have the intention
of replying publically (i.e. to the list) but your EMail
system may direct that message instead to the
indivdual, who may have posted the original question and
then the remainder of readers will
not see the information you are providing in response.
(Assuming of course that you wish public viewing which is
the purpose of this list).

Third, and lastly, on the question of resume posting.

I think we now have a concensus and let's call
the subject closed and get back to microscopy. The replies
most of which came to me directly (instead of the list)
are definitely in favor of posting announcements (~} 95%)
of position's/jobs and overwhealming opposed to resume's
(} 90%). I've also updated the welcome notice to make
this clear too.



-----------------------------------------------------
Nestor J. Zaluzec
Electron Microscopy Center For Materials Research
Argonne National Laboratory, Argonne, Ill. USA

Tel: 708-252-5075/4964 Fax: 708-252-4798
Email: Zaluzec-at-anlemc.msd.anl.gov
-----------------------------------------------------






From: Tony Hollenkamp :      afh-at-dmp.csiro.au
Date: Fri, 8 Oct 1993 11:48:47 +1000 (EST)
Subject: stm purchase

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Further to HJOAN's query on stm purchase, our Park Autoprobe is great
(i.e., fast and relatively inexpensive) for changing between techniques, iff
you want to do things other than stm.

Tony Hollenkamp






From: Jouko K. M„ki :      jokamaki-at-utu.fi
Date: Fri, 8 Oct 1993 09:44:11 +0200
Subject: Re: JoAn' s question about STM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Warning: Original message contained 8-bit characters, however during
the SMTP transport session the receiving system was unable to
announce capability of receiving 8-bit SMTP (RFC 1425-1428),
and as this message does not have MIME headers (RFC 1341) to
enable encoding change, only option was to STRIP sent characters
to 7-bit :-(



} On Thu, 7 Oct 1993 15:57:00 +0200, { HJOAN%CLEMSON.CLEMSON.EDU-at-ANLVM.CTD.ANL.GOV} wrote:
}
} }
} } resume's - Maybe in the future, this is just starting we need to give it a chanc
} } e. Question. We are in the process of purchasing STM. What are the important fea
} } tures to look for? Any information would be very much appreciated. Thank you for
} } your time, JoAn
}
Hello JoAn,

Based on experience of one year using Nanoscope II, I would say that one of
the most important features is the easy of localisation of the sample area.
It can never be very easy but many things can help you. There should be a
proper light microscope throug which you should be able to observe both the
specimen and the tip simultaneously perpendicular to the sample. This
means, you should have a reverse microscope and a stereomicroscope fitted
to the STM/AFM. Furthermore, the stereomicroscope should easily be turned
so, that you can observe the sample and the tip from the side.

These were on the top of my memory. In case there is more in my FILO-
memory, I will let you know.

Best regards,
Jouko

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Jouko K. Mki Navigare necesse est...
Laboratory Manager
Laboratory of Electron Microscopy
University of Turku Kiinamyllynkatu 10 FIN 20520 TURKU
INTERNET: jouko.maki-at-utu.fi Tel.: + 358 21 633 7318




From: Zaluzec-at-anlemc.msd.anl.gov (Nestor J. Zaluzec)
Date: Fri, 8 Oct 1993 09:44:11 +0200
Subject: AFM -spectral power density

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: Zaluzec-at-anlemc.msd.anl.gov (Unverified)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: microscopy-at-anlemc.msd.anl.gov



On Oct. 6th AG wrote:

} Anybody have access to power spectral density calculation programs for
} atomic force microscopy data? My AFM vendor is slow to provide this
} function, and I need it to help understand the "wavelength" of roughness
} on silicon surfaces. Any help appreciated!

} Andy Gilicinski


Can someone briefly explain "power spectral density calculations" in
this context and how it relates to surface roughness? I'm sure
I do not know the answer, but am interested in the question, what
does the calculation do? Is it a fourier transform? If so there
are plenty out in the public domain.


-----------------------------------------------------
Nestor J. Zaluzec
Electron Microscopy Center For Materials Research
Argonne National Laboratory, Argonne, Ill. USA

Tel: 708-252-5075/4964 Fax: 708-252-4798
Email: Zaluzec-at-anlemc.msd.anl.gov
-----------------------------------------------------






From: ROSEANN CSENCSITS (708) 252-4977, -7902 :      CSENCSITS-at-anlemc.msd.anl.gov
Date: Fri, 8 Oct 1993 13:17:59 -0500 (CDT)
Subject: pt to pt resolution tests

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For a 400kV microscope what is the amorphous material of choice for
point to point resolution tests? Carbon or Germanium?
I am using an optical bench to determine resolution.

Roseann Csencsits
Argonne Nat Lab
708-252-4977




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 8 Oct 1993 14:16:31 -0500 (CDT)
Subject: PD Software for EDS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



S. Bunasawa asks:
Is there a public domain or shareware EDS out there for the PC? I'm looking
for books and some reference so I can write my own EDS. Can someone
recommend a good source?


Reply, Send an Email message to EMMPDL-at-ANLEMC.MSD.ANL.GOV
in the message say "SEND HELP"
you will get instructions on how to access the EMMPDL by Email.
FTP will not be up for few weeks (minimum).

There are EDS programs in the XEDS directory in that library.

Nestor Zaluzec
ANL EMCenter




From: MOSSANT-at-DUCVAX.AUBURN.EDU
Date: Sun, 10 Oct 1993 11:05 CST
Subject: Primera printer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: Microscopy-at-anlemc.msd.anl.gov

Message to Jeff Ingeman:

Who sells this printer? Do you have the address of the manufacturer?
Thanks in advance.

tony Moss
Auburn University




From: VEI011-at-GEEL.DWT.CSIRO.AU (Colin Veitch CSIRO Division of Wool Technology)
Date: Tue, 12 Oct 1993 8:17:26 +1000 (EST)
Subject: NIST DTSA Program

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Some time ago the NIST published DTSA (Desk Top Spectrum Analyser) for Mac
systems. Does anyone know if there is a PC version available (or
something similar) which can accept either EMSA format files or Noran
Voyager format files. Is there something at the EMMPDL address (given
yesterday) which may be of use? If so, is that address an open one?

Thanks,

Colin Veitch





From: VEI011-at-GEEL.DWT.CSIRO.AU (Colin Veitch CSIRO Division of Wool Technology)
Date: Tue, 12 Oct 1993 11:19:43 +1000 (EST)
Subject: NIST/DTSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Alan,

NIST/DTSA is available through:
Office of Standard Reference Data 221/A323
National Institute of Standards and Technology
Gaithersburg, MD, 20899 USA
Tel (301) 975 2208
Fax (301) 975 0416

It was written by Chuck Fiori, Carol Swyt and Robert Myklebust.

This information is as printed on the manual and may have changed in the
past year or so. I have no idea regarding the price! 8-)

If you want more info. give me a call.

Colin Veitch

#####################################################################
*******************************
* Intuition is reason in a *
0------* hurry. *
} ---|--- { * H. Jackson *
| *******************************
/ \
_/ \_

Colin Veitch Tel + 61 (0)52 47 2611
CSIRO Division of Wool Technology Tel + 61 (0)52 47 2891 (dir.)
P.O. Box 21 Fax + 61 (0)52 47 2657
BELMONT Vic 3216
Australia

#####################################################################






From: K. M. Anisur Rahman :      anis-at-moe.eece.mu.edu
Date: Sat, 9 Oct 1993 20:38:12 -0500 (CDT)
Subject: How to identify these nano-crystals.. please help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hallo everybody:

We have taken some HREM micrograph of a composite ceramic
material, made with a crystalline phase dispersed into an amorphous
glassy phase. In the glassy phase there are small crystallites whose
diameter is of the order of few nanometers. We would like to identify
these nano-crystals. By measuring their lattice spacing no conclusive
identity can be assigned. It is also almost impossible to do an EDS on
such a small area.

We would appreciate very much if somebody can give us some
idea regarding how to identify these crystallites with certain degree of
confidence. Thank you.

-Anis.





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 11 Oct 1993 20:36:20 -0500 (CDT)
Subject: DTSA Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Reply to last 2 questions on DTSA.

At the presents DTSA only runs on a Mac platform. I do not
know if any plans are being made to expand it to PC.

The cost of DTSA is ~ $900-1000 US (as of a few months ago).
and it can be purchased directly from NIST. Demo copies are available
for the Mac platform. I do not have the address handy here at home
I will post the information tomorrow from the lab.

One of the priniciple authors of the program Chuck Fiori died
suddenly earlier this year. The program is still being supported
and presumably updated by colleagues at NIST, however, I'm not sure if
they wish their Email address to be given out. I will contact them
and ask them. If they concur I will post it on the listserver.

The EMMPDL has PC based programs for EDS data reduction, however,
they are not as comprehensive as DTSA. The EMMPDL is an open library,
you may receive information & source code by EMail to EMMPDL-at-ANLEMC.MSD.ANL.GOV
in your message include the single line "SEND HELP". Internet & FTP
access will be operating later this year (when I get all the bugs fixed!).
You may also login by telephone (however I see that the questions come
from colleagues in Australia), or you can make copies of the library at
you next local/national (?) EM society meeting.

Complete copies of the EMMPDL will be available at the next
Australian meeting in Brisbane, ICEM in Paris, and MSA in New Orleans
and any other EM Society meeting that makes arrangement with me in
advance.

=================
Nestor J. Zaluzec
ANL EM Center





From: Colin MacRae :      cmac-at-dmp.csiro.au
Date: Mon, 11 Oct 1993 13:51:37 +1000 (EST)
Subject: Re Joans question regarding STM/AFM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


MAIL FROM: {cmac-at-dmp.csiro.au}
RCPT TO: {MICROSCOPYLIST}
ARRIVAL_TIME: 11-OCT-1993 07:46:57


JOAN,
I have recently bought a PARK AutoProbe LS AFM/STM with
Electrochemical AFM and STM. The list of requirements when considering
buying such instruments is quite long.
Briefly
consider what sample size you wish to scan,
what detail do you want to see,
what modes of operation, AFM, STM, ECAFM, ECSTM, LFM, MFM etc,
what optical resolution do you need to locate your features of interest,
are you happy with a windows look a like interface or do you want a true
Windows interface.

If you approach PARK they will supply you with a booklet "How to Buy a
Scanning Probe Microscope". It will be quite helpful. Try to visit SPM
labs to see how samples are prepared and imaged. As a matter of interest
the ACEM-13 conference to be held in Queensland will have a one day
workshop on SPM. As I am one of the workshop organisers I think it would
be quite useful. However, a long way to travel. I'm sure there must be
closer conferences.

Good luck

Colin MacRae
Electron Microscopy Section

Commonwealth Scientific and Industrial
Research Organisation. _--_|\ cmac-at-dmp.CSIRO.AU
Division of Mineral Products / \ +61 3 647 0296
PO Box 124, Port Melbourne 3207 \_.--._/
AUSTRALIA v





From: Matt Madison :      MADISON-at-FLAMIN-O.TGV.COM
Date: Tue, 12 Oct 1993 10:03:39 -0700 (PDT)
Subject: test

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


MAIL FROM: {jmichael-at-vnet.IBM.COM}
RCPT TO: {MICROSCOPYLIST}
ARRIVAL_TIME: 12-OCT-1993 08:44:53

Please ignore this test message. Thank you.




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 12 Oct 1993 12:37:34 -0500 (CDT)
Subject: Positron Annihilation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-anlemc.msd.anl.gov


The Positron Annihilation facility at Argonne National Lab
was shutdown about a year ago, when the funding was cut.
In a loose interpretation of "microscopy" as those techniques
which allow the experimentalist to determine the structure of
a material which cannot be analyzed by the naked eye, then
PAS fits. Remember we are not prejudice to any specific microscopy
in this forum...

Nestor Z.
ANL EMCenter




From: DUNLAP-at-utkvx.utk.edu
Date: Tue, 12 Oct 1993 08:32:09 -0400 (EDT)
Subject: Address for list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thank you for adding my name to the list server. My correct address is
DUNLAP-at-utkvx.utk.edu.
Best Wishes
John Dunlap
The University of Tennessee
Knoxville, TN 37996-0810




From: ROSEANN CSENCSITS (708) 252-4977, -7902 :      CSENCSITS-at-anlemc.msd.anl.gov
Date: Mon, 11 Oct 1993 15:14:08 -0500 (CDT)
Subject: pt to pt resolution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Last week I asked:
For a 400kV microscope what is the amorphous material of choice for
point to point resolution tests? Carbon or Germanium?

Most answers were that Germanium was preferred and the second choice
was Silicon--because it is readily available in most labs.

Thanks for the confirmation.

Roseann




From: K. M. Anisur Rahman :      anis-at-moe.eece.mu.edu
Date: Sat, 9 Oct 1993 20:38:12 -0500 (CDT)
Subject: How to identify these nano-crystals.. please help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


MAIL FROM: {jmichael-at-vnet.IBM.COM}
RCPT TO: {MICROSCOPYLIST}
ARRIVAL_TIME: 12-OCT-1993 08:44:53

Hallo everybody:

We have taken some HREM micrograph of a composite ceramic
material, made with a crystalline phase dispersed into an amorphous
glassy phase. In the glassy phase there are small crystallites whose
diameter is of the order of few nanometers. We would like to identify
these nano-crystals. By measuring their lattice spacing no conclusive
identity can be assigned. It is also almost impossible to do an EDS on
such a small area.

We would appreciate very much if somebody can give us some
idea regarding how to identify these crystallites with certain degree of
confidence. Thank you.

-Anis.





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 12 Oct 1993 14:06:32 -0500 (CDT)
Subject: Multiple Messages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


October 12, 1993

All Microscsopy Subscribers

The mailserver system had locked up on a series of
addresses to Germany, resulting in a huge queue of
possible lost/deleted/replicatd mail.

You may be already aware of this as you may be
seeing duplicates of mail you already have received.
This should only take a day to clear up. There
was a queue of about 100 messages here at this end,
some fraction of which were destined for the Microscopy
server. The rest to users of the EM Center here at Argonne.
Sorry for the duplicate traffic, but I was not able to spend
the time necessary to read each of these messages and then
sorting out to whom it belonged and if it had already
been sent and the simplest method was just to restart
the system and let it clear itself out.

Hopefully ? :-\ this won't happen again.

Cheers- Nestor
ANL EM Center





From: llsutter-at-mtu.edu
Date: Tue, 12 Oct 1993 16:11:03 -0400
Subject: Multiple Messages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I'm not sure what the cost of a CCD camera is but I would guess
that it is near the cost of a low-light video camera and control box from
Custom Camera in the UK. The majority of the excess cost in once
commercially available systems was going to the vendor, not the
manufacturer of the camera. The advantage of the low-light video camera
and control box is that the pattern is instantaneous. My understanding of
the CCD camera is that it takes a period of time to accumulate each
pattern. This makes simple tasks such as comparing adjacent grains or
aligning a specimen more time consuming.

Thanks



Larry Sutter
Scattering Events R' Us





From: mecavaleri-at-mmm.com
Date: Tue, 12 Oct 1993 16:38:03 -0500
Subject: Listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is the fastest I have ever started receiving a subscription




Mark E. Cavaleri
3M CRL/A&PRL Microscopy Laboratories
3M Cneter 201-1E-15
3M Center 201-1E-15
St. Paul, MN 55144-1000
(612) 733-3247
(612) 733-0648 FAX




From: llsutter-at-mtu.edu
Date: Tue, 12 Oct 1993 16:32:24 -0400
Subject: Bob Keller's EBSP Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I'm not sure what the cost of a CCD camera is but I would guess
that it is near the cost of a low-light video camera and control box fom
Custom Camera in the UK. The majority of the excess cost in once
commercially available systems was going to the vendor, not the
manufacturer of the camera. The advantage of the low-light video camera
and control box is that the pattern is instantaneous. My understanding of
the CCD camera is that it takes a period of time to accumulate each
pattern. This makes simple tasks such as comparing adjacent grains or
aligning a specimen more time consuming.

Thanks



Larry Sutter
Scattering Events R' Us





From: Colin MacRae :      cmac-at-dmp.csiro.au
Date: Mon, 11 Oct 1993 13:51:37 +1000 (EST)
Subject: Re Joans question regarding STM/AFM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



JOAN,
I have recently bought a PARK AutoProbe LS AFM/STM with
Electrochemical AFM and STM. The list of requirements when considering
buying such instruments is quite long.
Briefly
consider what sample size you wish to scan,
what detail do you want to see,
what modes of operation, AFM, STM, ECAFM, ECSTM, LFM, MFM etc,
what optical resolution do you need to locate your features of interest,
are you happy with a windows look a like interface or do you want a true
Windows interface.

If you approach PARK they will supply you with a booklet "How to Buy a
Scanning Probe Microscope". It will be quite helpful. Try to visit SPM
labs to see how samples are prepared and imaged. As a matter of interest
the ACEM-13 conference to be held in Queensland will have a one day
workshop on SPM. As I am one of the workshop organisers I think it would
be quite useful. However, a long way to travel. I'm sure there must be
closer conferences.

Good luck

Colin MacRae
Electron Microscopy Section

Commonwealth Scientific and Industrial
Research Organisation. _--_|\ cmac-at-dmp.CSIRO.AU
Division of Mineral Products / \ +61 3 647 0296
PO Box 124, Port Melbourne 3207 \_.--._/
AUSTRALIA v






From: llsutter-at-mtu.edu
Date: Tue, 12 Oct 1993 21:10:54 -0400
Subject: EBSP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310071557.AA18638-at-enet-gw.pa.dec.com}


I forgot to mention that I have captured EBSP patterns on a Mac
and imported them into the Image 1.47 program. I think the macro language
of the program could be used to index the patterns but have never had time
to try. Has anyone else? If not, the code of the program could be
readily modified to do the job. Has anyone explored this option?

Thanks
llsutter-at-mtu.edu
Larry Sutter
Scattering Events R' Us





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 12 Oct 1993 23:05:50 -0500 (CDT)
Subject: House Cleaning Done

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


MAIL FROM: {beanland-at-liverpool.ac.uk}
RCPT TO: {MICROSCOPYLIST}
ARRIVAL_TIME: 12-OCT-1993 06:43:27


Microscopy Subscribers

The purge of old message files is done, and the
duplicate messages should be gone. I believe the
software update to Mutlinet is now complete and
things should be back to normal. Thanks for
your patience.

Nestor




From: llsutter-at-mtu.edu
Date: Tue, 12 Oct 1993 21:10:54 -0400
Subject: EBSP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"Microscopy Mailing List" {microscopy-at-anlemc.msd.anl.gov}

Reply_ Re: EBSP
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
A student here has modified NIH-Image 1.45 to do some basic analysis of EBSPs.
We capture EBSPs directly into Image with a Scion LG-3 frame grabber and
Dage-MTI camera. The camera points at a phosphor screen through the viewport
of the SEM (we actually use an ESEM - nice, plenty of room and no worries about
light leaks!!). We have a paper explaining the design and construction almost
ready for submission. Preprints will be available on request. The program
modifications are a currently little cryptic (i.e. no manual) and only work on
cubic crystals but they do allow you to do orientation analysis. The output is
further analyzed by programs called MisMat and FindCSL which are both in both
the EMMPDL and Microbeam Analysis Society Software Library. They are Mac
programs. I will ask the student if he will release his Image code early if
people are interested. If he does, no complaints please, he's not a
programmer, just a Nuc Eng student trying to get a Ph.D! Modofocations are in
Pascal not the built in macro language.
The other programs I mentioned above are available on
freebie.engin.umich.edu in the dir /pub/MSA+MAS/
both the EMMPDL and MASSL are available there for anonymous ftp.
If you have difficulty connecting or downloading any software from this site
send mail to me.

OK?
John Mansfield
--------------------------------------

I forgot to mention that I have captured EBSP patterns on a Mac
and imported them into the Image 1.47 program. I think the macro language
of the program could be used to index the patterns but have never had time
to try. Has anyone else? If not, the code of the program could be
readily modified to do the job. Has anyone explored this option?

Thanks
llsutter-at-mtu.edu
Larry Sutter
Scattering Events R' Us


------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;12 Oct 1993 21:56:51 U
Received: by totalrecall.rs.itd.umich.edu (5.67/2.2)
with X.500 id AA25695; Tue, 12 Oct 93 21:54:33 -0400
with SMTP id AA25690; Tue, 12 Oct 93 21:54:31 -0400








From: llsutter-at-mtu.edu
Date: Tue, 12 Oct 1993 16:11:03 -0400
Subject: CCD for EBSP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310131449.AA08908-at-riker.ml.wpafb.af.mil}
"Microscopy Mailing List" {microscopy-at-anlemc.msd.anl.gov}

Reply_ CCD for EBSP
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
}
} I'm not sure what the cost of a CCD camera is but I would guess
} that it is near the cost of a low-light video camera and control box
} from Custom Camera in the UK. The majority of the excess cost in once
} commercially available systems was going to the vendor, not the
} manufacturer of the camera. The advantage of the low-light video camera
} and control box is that the pattern is instantaneous. My understanding
} of the CCD camera is that it takes a period of time to accumulate each
} pattern. This makes simple tasks such as comparing adjacent grains or
} aligning a specimen more time consuming.
}
} Thanks
}
}
}
} Larry Sutter
} Scattering Events R' Us

Not really true! Both cost wise and time wise.
A slow scan camera would be exepnsive but we use a TV rate Camera (A Dage-MTI
CCD72) and a cheap frame capture board (Scion LG3) to record our EBSPs.
We use the video rate capture of the LG3 to grab several consecutive frames of
video and average those. It is obviously not as noise free as a good slow
scann camera but our EBSPsystem cost less than $12K including the Macintosh!
Works nicely too.

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;12 Oct 1993 17:48:50 U
Received: by totalrecall.rs.itd.umich.edu (5.67/2.2)
with X.500 id AA17248; Tue, 12 Oct 93 17:46:30 -0400
with SMTP id AA17243; Tue, 12 Oct 93 17:46:27 -0400








From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Wed, 13 Oct 1993 08:18:34 -0700 (PDT)
Subject: Re: EBSP systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-carson

How does the 6100 compare to the 6300F? The 6300 has a BNC port that is
live from the monitor. We wheeled a MacII cx outfitted with a Data Transl
ation frame grabber and NIH Image into the scope room. The DT board was
connected with a coax cable to the bottom-most BNC connector on circuit
board U-34. Get to this board by removing the kick panel below the
operator's console. Look for a series of boards to your left. You have
to time the capture to avoid getting a scan shift if you have any image
drift - simply issue the command to grab as the scan gets to the bottom.
You can't use the SR mode with this. If you can access the video on the
610, then you can get near publication quality images with Image. I even
got some 4X5 publication quality prints by helping the contrast with Adobe
Photoshop and printing it out on our campus print shop's Linotronic press.
Of course you might try calling JEOL technical folks.
On Wed, 6 Oct 1993, Bob Keller wrote:

} We would like to set up a simple system for recording electron backscattering
} patterns in a JEOL 6100, without resorting to some of the (overpriced)
} commercial systems currently available. At the moment, we are considering the
} use of a phosphor screen and CCD set-up, using an optical viewing port in the
} SEM. As the patterns would be used simply for orienting specimens for the time
} being, publication-quality images are not necessary. We do have access to
} image processing via NIH-Image for contrast enhancement/noise reduction after
} acquiring images. Has anyone tried doing this with success? Without success?
}
} Bob Keller
} NIST-Boulder
}







From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 13 Oct 1993 11:10:18 -0500 (CDT)
Subject: DTSA Information

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Microscopy Subscribers:

As promised here I checked with NIST and here is the information
about the DTSA (Desk Top Spectrum Analyzer) Program for
EDS data analysis.

ORDERING INFORMATION:
contact Joan Sauerwein Email: srdata-at-enh.nist.gov

Technical Questions:
contact Bob Myklebust Email: myklebust-at-gapnet.nist.gov

General questions can always be posted here at the Microscopy
server since the NIST group is one of our subscribers. By posting
some of the questions here you may receive additional comments
insight and/or answers from other users on this server. Remember
we will all benefit from this type of discussion in the long
run!


Nestor Zaluzec
ANL EM Center




From: ROSEANN CSENCSITS (708) 252-4977, -7902 :      CSENCSITS-at-anlemc.msd.anl.gov
Date: Thu, 14 Oct 1993 17:28:20 -0500 (CDT)
Subject: beam energy spread

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


HI

Has anyone measured the energy spread of a 400kV beam in a JEOL 4000
with a saturated or undersaturated LaB6 filament? Often we assume
a 1 ev energy spread but is this reasonable or is it more like 3-4ev
energy spread?

Thanks for your help.

Roseann Csencsits
Electron Microscopy Center-Argonne Nat. Lab




From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Thu, 14 Oct 1993 18:58:38 -0700 (PDT)
Subject: JEOL 6100 and Image

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-carson.u.washington.edu

} How does the JEOL 6100 compare to the 6300F? The 6300 has a BNC port
that is
live from the monitor. We wheeled a MacII cx outfitted with a Data
Translation frame grabber and NIH Image into the scope room. The DT
board was connected with a coax cable to the bottom-most BNC connector
on circuit board U-34. Get to this board by removing the kick panel
below the operator's console. Look for a series of boards to your
left. You have to time the capture to avoid getting a scan shift if
you have any image drift - simply issue the command to grab as the
scan gets to the bottom. You can't use the SR mode with this. If you
can access the video on the 6100, then you can get near publication
quality images with Image. I even got some 4X5 publication quality
prints by helping the contrast with Adobe Photoshop and printing it
out on our campus print shop's Linotronic press. I was tipped off about
the presence of this port while speacking to our local JEOL rep. about
getting my hands on the images stored on the 6300's internal harddrive.
The stored images were going to require about $20,000 to get at since
many years ago JEOL adopted an obscure and difficult (their words) image
storage format.

I've tried to post this several times and our mail
server has been having problems. If this got through earlier, sorry for
adding to the chatter.







From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Thu, 14 Oct 1993 21:40:01 -0500 (CDT)
Subject: Contamination

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Your dark "contamination" spots are "cracked" hydrocarbons from the
a variety of sources. 1.) Diffusion Pump Oils, 2.) Surface contamination
(fingerprints etc...), 3.) Electropolishing solutions.... They are attracted
electrostatically to the the beam perimeter (small probes are worse than
large probes) and essentially get baked
on to your specimen. The phenomenon occurs in all microscopes I've
used (and I've used a few), however, it can be minimize depending
upon it's source. If you have a poor vaccum ( ~ 10-7 Torr)
then cool your specimen to ~ -30 C. The residual H20 vapor in your
scope will begine to collect near your specimen surface,
the action of the electron beam will
dissociate the H & O, the O will react with the C to form CO2 gas
and the contamination will cease. Of course if your specimen is
carbon based then you drill holes (I did a nice job on a diamond
specimen once and learned the hard way). You can also heat your specimen
but this doesn't always work as well. Spreading the beam to cover a
large area and heating the local region (remove all apertures and
go to the largest spot size) will temporarily immobilize things but
eventually (after about 1/2-1 hour) things will begin to diffuse
across your specimen to the beam area again. There are a variety of other
tricks which depend on your specimen. For example semiconductors
contaminate less than metal (in clean vaccum system { 10-9 T)
A starting reference you should look up is
Introduction to Analytical Electron Microscopy, (1st edition not
the 2nd) Ed. Hren, Goldstein, Joy, 1979, Plenum Press,
look at Chapter 18, by John Hren
entitled "Barriers to AEM: Contamination and Etching".....


Nestor Zaluzec
ANL EM Center




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Thu, 14 Oct 1993 21:42:55 -0500 (CDT)
Subject: Contaminination Continued

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Oh yes it appears dark because it is so thick!. The spots can
build up to heights of microns. You can easily see this by tilting
your specimen 30-45 degrees after the spots form. The nice round
spots become tilted cones......


Nestor





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 15 Oct 1993 10:18:19 -0500 (CDT)
Subject: STEM probe size

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



It is bad policy to believe any manufacturers claim on the probe
size. It will be a function of C2 aperture size, lens excitation,
specimen height in the lens and a variety of other parameters. You
can get a quick and dirty measurement by doing a line scan across
a sharp edge and looking at the signal rise profile when going from
a "hole" to a specimen, however, you will need a calibrated magnification
to do this. Another bad policy of the manufacturers is that they
will claim the probe size based upon FWHM measurements, you should
of course be using at least FWTM or even better the first zero of the
crossing of the interference function.

A better way to measure the probe size is to image the STEM probe
in the TEM mode. Can you turn on your post-specimen lenses and
configure them to image the probe in the JEOL 2000? This is possible
in some microscopes, but I'm not sure about the 2000 series...


Nestor Z.
ANL EM Center




From: Paul Goodwin :      pgood-at-inson.fhcrc.org
Date: Fri, 15 Oct 1993 09:35:43 -0700 (PDT)
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopy





From: {QJXNJ21-at-memrqa.sps.mot.com}:ddn:wpafb
Date: 10-15-93 11:01am
Subject: GRAIN SIZE DISTRIBUTION MEASUREMENT

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310151659.AA15024-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb
Subj: GRAIN SIZE DISTRIBUTION MEASUREMENT
Orig-Author: {QJXNJ21-at-memrqa.sps.mot.com}:ddn:wpafb
-----------------------------------------------------------
Hello readers,
Thanks to all who responded to my question about contamination spots.
This IS a great format to have such questions answered. Now I have
another problem which I think many TEMmers must have been up against.
How to measure grain size distributions from TEM images of polycrys-
talline materials e.g. polySi or Al etc. What kind of image proces-
sing algorithms should be used and with what strategy? I realize that
a generalized algorithm may not work for every image but there must
be some general principles. I've heard that the Laplacian filter
shows up edges but I also heard John Russ of NCSU say how he uses
repetitive filtering and subtraction procedures to achieve grain
boundary delineation. It is not intuitively apparent to me how
this multiple filtering is used. Any tips or insights will be
highly appreciated. We have Gatan's Digital Micrograph and the
NIH Image 1.4 software. I dont want to use the crude straight
line method because I want to calculate size distributions.
Thanks for your time.
Vidya Kaushik
TEM Facility, MOTOROLA INC.,
Austin, TX, 78721 (512)-928-5134; (512)-928-6664 fax
QJXNJ21-at-MEMRQA.SPS.MOT.C





From: Paul Goodwin :      pgood-at-inson.fhcrc.org
Date: Fri, 15 Oct 1993 10:26:58 -0700 (PDT)
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



unsubscribe Paul Goodwin






From: mecavaleri-at-mmm.com
Date: Fri, 15 Oct 1993 12:53:34 -0500
Subject: Confocal Microscopy in Industry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310151739.AA01127-at-pascal.mrd.bldrdoc.gov}

I would like to find out how laser scanning confocal microscopy
is being used industry (other than semiconductors). Does this microscopy
really provide a significant improvement over 'ordinary' light microscopy?

Mark E. Cavaleri
3M CRL/A&PRL Light Microscopy
3M Center 201-1E-15
St. Paul, MN 55144-1000
(612) 733-3247
(612) 733-0648 FAX




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 15 Oct 1993 13:37:14 -0500 (CDT)
Subject: Probe size Measurements

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Using the signal rise profile technique to get a nominal probe
size is valid for both convergent & parallel probes, whether
you are in STEM, TEM, SEM or any other mode..... It's major
draw back is that it is not sensitive to weak low intensity
tails which can cause signal generation in modes other than
CBED, for example, as in X-ray analysis.

Nestor Z.
ANL EM Center




From: mecavaleri-at-mmm.com
Date: Fri, 15 Oct 1993 15:44:53 -0500
Subject: re-send of Confocal query

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to find out how laser scanning confocal microscopy
is being used industry (other than semiconductors). Does this microscopy
really provide a significant improvement over 'ordinary' light microscopy?

Mark E. Cavaleri
3M CRL/A&PRL Light Microscopy
3M Center 201-1E-15
St. Paul, MN 55144-1000
(612) 733-3247
(612) 733-0648 FAX




From: jcrum%sunstroke%sdsu.edu-at-relay.tc.umn.edu (John Crum)
Date: Fri, 15 Oct 1993 18:38:14 -0500 (CDT)
Subject: SEM Image Capture

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Microscopy Subscribers:



NIH Image Users:

Is anyone out there using NIH Image and a Macintosh to capture images from
a scanning electron microscope? If so, what is the resolution (i.e. image
size) that you are getting? What hardware (i.e. image capture board and
Mac) are you using? What does it cost?

The reason I ask is that our EM facility is in the market for a new SEM.
A nice feature we have now on our SEM is image capture, using a part of a
Princeton Gamma Tech EDS system. The core of that system, though, is a
Sun 3/60, which doesn't really work well and will need a costly upgrade.
We can capture images of 512 pixels and 1024 pixels sqaure. New software
(on top of hardware upgrade) will capture 2kX2k images.

So, the ultimate question is: should we adapt the system we have now to a
new SEM, or buy a new Mac/Image based system?

Any advice is greatly appreciated.

jc





From: smithj-at-acad.winthrop.edu
Date: Mon, 18 Oct 1993 09:15:31 -0400
Subject: Video for TEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
I'm interested in hooking a video camera to my TEM (JEOL 100B):
1)To demonstrate procedures such as filament saturation and 'scope
alignment when teaching my EM class, and 2) To capture "quickie"
EM prints to a video printer for students in my cell bio and vertebrate
histology classes. I'm looking for about 640x480 pixels and 8 bits, and
would like something that covers about the same area at each
magnification as the film camera.
I know about Gatan's systems, and I've been told "look at the Fullam
system before you buy it". Is there anything else out there that I should
know about? Have any of you used either the Gatan system or the Fullam
system, and how satisfied are you with it? Any help/feedback would be
appreciated.
TIA
Julian Smith III
(smithj-at-winthrop.edu)




From: :      tonygr-at-EAGLE.MIT.EDU
Date: 18 Oct 93 08:33:32 EDT
Subject: Video on TEMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9310181433.AA27767-at-EAGLE.MIT.EDU}

-------------

This is in response to Julian Smith's query about video systems on TEM's.

I only have experience with the old Gatan system, so can't really comment
on the questions directly. There is an issue, however, with just taking
'Quickie video prints'. A video printer captures only a single frame of the
RS170 video - that is, it is equivalent to a 1/30 sec. exposure. During
this time, one can, (depending on the microscope operating conditions) get
significant shot noise if the video system is reading out the electron
image in real time (as does the old Gatan image intensifier system). Such
a print can be next to useless. Since our eyes respond much more slowly,
and in any case we are viewing a phosphor (either the microscope screen
or the screen of the CRT) which adds its own time constant, we don't notice
the effect so much (although it is resonsible for the "liveliness" one sees
when viewing a high magnification image on the viewing screen).

This is not an issue when reading out a framestore loaded from a slow-scan
system, but they are expensive for just demonstrating saturation and taking
quickie video prints. The Fullam system is a light camera viewing a
phosphor screen - you will have to experiment to see if it has enough time
constant for your needs.

The RS170 can be frame-averaged either by a stand-alone system such as is
sold by Gatan or Synoptics (not terribly expensive) or by a board from one
of many vendors which plugs into a PC. The advantage of the latter, besides
being even less expensive, is that it also can transfer the image to the
PC for networking, image processing, etc.

Hope this helps.

Tony Garratt-Reed







From: llsutter-at-mtu.edu
Date: Mon, 18 Oct 1993 12:15:31 -0400
Subject: Stand Alone Correction Programs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I am looking for a stand-alone ZAF program that I can run on a PC
or Mac that will allow me to import k-ratios or app. concentration. I
specificly want to do carbonate analysis. A stand alone Bence-Albee
program would also suffice. Any thoughts...

Thanks
llsutter-at-mtu.edu
Larry Sutter






From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 18 Oct 1993 11:27:11 -0500 (CDT)
Subject: Destructions on Sub/Unsub

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



As requested here's a brief synopsis of the Subscribe/Unsubscribe
commands

**************************************************
* IMPORTANT: *
* please send this reply to *
* LISTSERVER-at-ANLEMC.MSD.ANL.GOV *
*------------------------------------------------*
* DONOT send it to Microscopy-at-anlemc.msd.anl.gov *
* otherwise everyone on the server will *
* see your message and this creates unnecessary *
* and annoying traffic!!!!!!!!!!!! *
**************************************************

Help can be obtained by sending an Email request to

Listserver-at-ANLEMC.MSD.ANL.GOV

in the Email request you should include one of the
following command lines (one command/line)

Command: Send Help
Result: Sends this file

Command: Subscribe {list} Username-at-EmailAddress
Result: Adds Username-at-EmailAddress to the list
Example: Subscribe Microscopy Zaluzec-at-anlemc.msd.anl.gov

Current {lists} supported:
{list}
Microscopy


Command: Unsubscribe {list} Username-at-EmailAddress
Result: Removes Username-at-EmailAddress the specified {list} .
Example: Unsubscribe Microscopy Zaluzec-at-anlemc.msd.anl.gov

=======================================================================
All other questions can be also sent as a text message to LISTSERVER
someone eventually reads all of them, but not at any particuliar time.
=======================================================================

=================
End of Message
=================




From: Denis Baskin :      baskindg-at-u.washington.edu
Date: Mon, 18 Oct 1993 17:23:44 -0700 (PDT)
Subject: cooled ccd cameras

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {23101815244205-at-vms2.macc.wisc.edu}

I am interested in using a cooled ccd camera for imaging fluorescence
immunostaining. I understand that there are advantages to the cooled
ccd-fluorescence microscope combination over a confocal microscope. I
would appreciate any information about cooled ccd camera-microscope
configurations from anyone who is (has) using one. What is the best
manufacturer? What microscope configuration is best? What about software?

Denis Baskin, Ph.D.
VA Medical Center and University of Washington
Seattle, WA
(206) 764-2138 FAX (206) 764-2164






From: jeanne_barker-at-merck.com
Date: 19 Oct 1993 08:33:42 U
Subject: Acid phosphatase

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subject: Time:8:24 AM
OFFICE MEMO Acid phosphatase Date:10/19/93
Can anyone out there recommend a protocol for acid phosphatase staining of 1
micron, epon embedded kidney sections? I have tried a couple of kits that are
suppose to remove the epon and the fixatives, but the results have not really
been spectacular.

If there are any biological electron microscopists out there who could give me
a reference, I would really appreciate it.

Thanks, Jeanne
Jeanne_Barker-at-Merck.Com






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Tue, 19 Oct 1993 14:56:52 -0700 (PDT)
Subject: Re: SEM Image Capture

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For our JEOL 6300F we found that we could simply string a coax cable from
the Data Translation frame grabber board in a Mac to a live video port
hidden behind the kick panel below the operator's console. This allowed
us to capture the signal to the SEM monitor with NIH Image. The command
to grab had to be synched by eye so that the scan raster had reached the
screen bottom. The images were decent quality, and with some Photoshop
grayscale and brightness adjustments we obtained publication quality
prints on the campus Linotronic press.
We aren't using this for anything useful yet for lack of a Mac to leave in
the 'scope room.






From: Gerald Little :      ANGJL-at-medicine.newcastle.edu.au
Date: Wed, 20 Oct 1993 09:28:51 GMT+1000
Subject: N-CAM Antibody for immunocytochemistry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.931020092851.288-at-medicine.newcastle.edu.au}

G'day Microscopy readers,

Help! I have been searching for a monoclonal antibody, Neural Cell
Adhesion Molecule (N-CAM), which will react with tissue that has been
fixed in at least 4% paraformaldehyde, preferably also with
glutaraldehyde. We are using cryo-ultramicrotomy on rat neural
tissue for TEM localisation of specific epitopes, N-CAM being one of
them. So far, none of the companies I have approached (Sigma,
Amersham, Serotec, Boehringer-Mannheim, Dako, Sera-lab) have an N-CAM
antibody which will work with formaldehyde fixed tissue. Would
any one out there know of a company which would have this particular
antibody.
Thanks in anticipation.
Gerry Little.
Dr Gerald Little | Ph (049) 215618
The Neuroscience Group |
Discipline of Anatomy | Fax (049) 216903
Faculty of Medicine |
The University of Newcastle | Email ANGJL-at-medicine.newcastle.edu.au
Australia, 2308 |




From: dfa-at-rabbit.physiol.unimelb.edu.au (David Abbott)
Date: Wed, 20 Oct 1993 10:30:33 +1000 (EST)
Subject: Electron microscopy list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To date, there has been much activity on this list, mostly
related to electron microscopy in one form or another.
It seems to me this much traffic warrants a list of its
own, and so I'd like to suggest that just such a "splinter"
list is created.

My personal reason for suggesting this is that, not being
involved in electron microscopy myself, I simply don't have
the time to look through all the EM messages.

Of course, those interested in all topics could subscribe
to both lists and be no worse off than now.

-David.
- -
David Abbott | Email: dfa-at-physiol.unimelb.edu.au
School of Physics & Dept. of Physiology | Phone: +61-3-344-5465
The University of Melbourne, Australia. | Fax: Available on request.





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 19 Oct 1993 20:44:17 -0500 (CDT)
Subject: INTERNET BBS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



For those Microscopy Listserver user that were trying.
The Internet link to the MSA BBS system is back up and
running. There was a software bug which took the line
down for about a week. It appears that everything is back
to some degree of normality......

Nestor Z.
ANL EMCenter




From: waheeschen-at-dow.com (Bill Heeschen (517)636-4005 PMRC/ASL)
Date: Tue, 19 Oct 1993 22:23:40 -0400
Subject: Re: Electron microscopy list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199310200209.AA04201-at-na1.dow.com}

Folks:

I recognize David's concern about lots of electron vs light microscopy, but my
personal feeling is that a single microscopy list is still the best way to go.
In our lab, most of the EM folks end up doing a bit of LM and vice versa.
Watching everything go by keeps the mind open and active. Message deletion is
a command-D away and the need can usually be determined within the first
sentence. If it gets to the point where LM is as busy as EM, then perhaps we
should consider a split, but until then my vote is a single list.

The mailing list for NIH Image (subscribe NIH-Image yourname-at-your.address to
listserv-at-soils.umn.edu) includes a fair amount of general digital image
acquisition and analysis discussion, often dealing with LM issues.

Lastly, if you plan to be away from your electronic in-box for a few days and
don't want to wade through buckets-o-bits when you return, unsubscribe from
your lists for the time you're gone.

Regards,

Bill

==========================
Bill Heeschen / Analytical Sciences - Materials Characterization
1897-D Building / The Dow Chemical Company
Midland, MI 48667 U.S.A.
phone: (517)636-4005 fax: (517)636-5453
Email: waheeschen-at-dow.com
==========================




From: morilak-at-cmgm.stanford.edu (David Morilak)
Date: Tue, 19 Oct 1993 18:57:51 PDT
Subject: N-CAM Antibody for immunocytochemistry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In reply to Gerald Little's request for sources of N-CAM monoclonals, I am
consulting the MSRS catalog (it's in front of me now!), and have come up with
this list of potentials:

Chemicon (definitely have polyclonals, call about monoclonals)
BioDesign, clone NKINBL1, cat # M6127oM
Accurate, clone NCAM-OB11, cat # BYA-6075-1
Development Studies Hybridoma Bank, 5B8
Development Studies Hybridoma Bank, 4d
Development Studies Hybridoma Bank, 5e
Development Studies Hybridoma Bank, 5A5 (sialylated form)
Santa Cruz Biotechnology, clone ERIC1, cat # sc-106
American Qualex, cat # M1180
Bioprobe/Thamer (?), 123C3 (I am not familiar with this company)
British Biotechnology, clone ERIC-1, cat # BCA8
Becton-Dickinson, clone MY31, cat# 550009
CRB/ICI, clone ERIC-1, cat # AM-19-500

I hope this helps!

David Morilak
Dept Psychiatry
Stanford Univ
morilak-at-cmgm.stanford.edu

-------




From: Darcy Clark :      03249587-at-mama.minmet.uq.oz.au
Date: Wed, 20 Oct 1993 15:50:40 +1000 (EST)
Subject: Re: INTERNET BBS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Tue, 19 Oct 1993, Nestor J. Zaluzec (708)-252-5075, -4964 wrote:

}
} For those Microscopy Listserver user that were trying.
} The Internet link to the MSA BBS system is back up and
} running. There was a software bug which took the line
} down for about a week. It appears that everything is back
} to some degree of normality......
}
} Nestor Z.
} ANL EMCenter

Just a quick question - what is the MSA BBS and how does one gain access
to it ??
Darcy Clark
Uni. of QLD
Brisbane.AUSTRALIA







From: modb%BANRUC60.BITNET-at-ANLVM.CTD.ANL.GOV
Date: Wed, 20 Oct 1993 10:06:21 +0100
Subject: A lot of mail ???

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Do the subscribers of the list get that amount of mail every day ?
I didn't receive any messages between Oct 6 and today.
Did I miss something important or does the list allow hidden options (USA anly)

Marc




From: howelld-at-egr.msu.edu
Date: Wed, 20 Oct 1993 08:05:58 -0400 (EDT)
Subject: Multilayer TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear Readers,

I would like to initiate a discussion on the subject of cross-sectional
TEM of multilayers (esp. metal/a-silicon). I am interested
in the alternate techniques to HREM that have appeared
in the literature such as the High-Angle Annular
Dark Field (HAADF) technique practiced by Stearn's group at ASU
(ref. Cheng et al., 1992, Proc. 50th Ann. Meet. EMSA, p.122) and the
Fresnel Method employed by Stobb's group at Cambridge (ref. Shih and
Stobbs, 1990, Ultramicroscopy, 32, p. 219).

The existence of these techniques arose from the concern of the ability
of HREM to characterize the roughness of the interface between the
alternating material layers which comprise the multilayer. Since many
multilayers are composed of two materials with vastly different Z (e.g.
X-ray mirrors) the strong scattering at the interface may complicate
the interpretation of HREM images.

What I would like to know is why these two techniques are not more
widely used. Although all of the HAADF work I have seen reported has
been done on VG-STEMs, I have seen a report from the Philips Electron
Optics Lab that the technique can be done on CM12,20,&30 TEM/STEMs
with annular dark field detectors (Otten, 1991, J. Electron Microscopy
Tech., 17, p.221)

As for the Fresnel Method (fitting exp. through-focal fresnel-fringe
line-profile intensities to those based on mixed scattering potentials
at the interface) I have counted over 20 papers from the
Cambridge group but have not come across one from anywhere else. One
impediment is the need to digitize the fresnel-fringe profiles with
fairly high resolution, which can be a problem for those of us without
access to a microdensitometer or the ability to collect the images
digitally in the first place. But is the method acceptable on a
theoretical basis?

If anyone has had experience with or has critically evaluated either
technique I would like to hear from you.

Thank You,

David A. Howell | Ph (517)337-2943
Dept. of Materials Science & Mech. | Fax (517)353-9842
Michigan State University |
E. Lansing, MI 48824-1226 | Email howelld-at-egr.msu.edu




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 20 Oct 1993 8:45:16 -0500 (CDT)
Subject: Internet BBS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Just a quick question - what is the MSA BBS and how does one gain access
} to it ??


In addition to this Listserver the ANL EMC and MSA have several other
"on-line" so to speak services. One is the EMMPDL (the Microscopy
and Microanalysis Public Domain Library of Software) and the
Electronic Bulletin Board System. Access to information about the EMMPDL
and the BBS as well as their contents are available via conventional
telecommunications (modem lines) or over Internet (via Telnet, not
yet FTP , but it is coming).

The BBS is a more selective forum than this list having society information,
announcement, job placement, topical discussion session, public
and private Email..... that is the type of information which
(for the most part) doesnot belong on this Email list.

To get internet information about the EMMPDL send an Email
message to "EMMPDL-at-ANLEMC.MSD.ANL.GOV" with the message
Send Help EMMPDL

To get internet information about the BBS send an Email
message to "EMCBBS-at-ANLEMC.MSD.ANL.GOV" with the message
Send Help EMCBBS

----------------------

Nestor Zaluzec
ANL EMCenter




From: AB :      morilak-at-cmgm.stanford.edu
Date: Wed, 20 Oct 1993 10:41:35 PDT
Subject: MSRS catalog of primary antibodies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have received several questions regarding the MSRS catalog of Primary
Antibodies that I mentioned in a previous transmission, so I am sending this
to both the microscopy and NIH image distribution lists. The MSRS catalog is
indeed a catalog of primary antisera, both those available commercially and
those available from individual labs who choose to have their antisera listed.
I have found it to be an extremely useful first reference, though it is, as
might be expected, not 100% comprehensive. I often will call the sources
listed to find out what else they have available that is not in the catalog
(e.g. polyclonals vs monoclonals, different host species, different epitopes
etc.). I ordered it at last years Neuroscience meeting, and an update (I am
told) will be available approximately annually. The cost was somewhere around
$75 (US). (MSRS stands for "Manufacturer's Specifications and Reference
Synopsis"). I don't have a phone number, but here is the address off the order
form:


MSRS catalog of Primary Antibodies
AERIE Corporation
Box 1356
Birmingham, MI 48012-1356
-------




From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Wed, 20 Oct 1993 17:06:20 -0700 (PDT)
Subject: Re: MSRS catalog of primary antibodies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-carson.u.washington.edu



On Wed, 20 Oct 1993, AB wrote:

} I have received several questions regarding the MSRS catalog of Primary
} Antibodies that I mentioned in a previous transmission, so I am sending this
} to both the microscopy and NIH image distribution lists. The MSRS catalog is
} indeed a catalog of primary antisera, both those available commercially and
} those available from individual labs who choose to have their antisera listed.

How does the MSRS compare to Linscott's Directory?

Glen MacDonald
Hearing Development Laboratories
UW






From: LAURAN%NKI.NL-at-ANLVM.CTD.ANL.GOV
Date: Thu, 21 Oct 1993 10:44 +0000 (GMT)
Subject: Cytogenetics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Does anyone know whether there is a Email or NEWS group specifically
relating to cytogenetics (methodology etc.)?
If so, please drop me aline where to find them.

Lauran Oomen, Netherlands Cancer Institute, Amsterdam, The Netherlands





From: LAURAN%NKI.NL-at-ANLVM.CTD.ANL.GOV
Date: Thu, 21 Oct 1993 11:05 +0000 (GMT)
Subject: CCD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We consider buying a microscope-CCD set up, which will be used in a
multi-user environment. Applications will comprise a.o. detection of
(low levels of) immunofluorescent signals in tissues and cells, in situ
hybridization signals and very low levels of NADH fluorescence.
Precise quantification and a large dynamic range are also a main issue.
Any suggestions relating to brands and types for the several components
of the total system (microscope, CCD camera, computer hardware and
software) will be highly appreciated.

Lauran Oomen, Netherlands Cancer Institute, Amsterdam, The Netherlands





From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Fri, 22 Oct 1993 11:22:20 -0600
Subject: Measurements of surface layer thichness

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0oqOc6-0004YCC-at-apus.cus.cam.ac.uk}

I'm looking for a technique to measure the thickness of surface
contamination/oxidation on metal surfaces. I need an image of the surface
for mapping purposes, as well as thickness measurements of the
contamination. These specimens range in size from 1-5 cm and less than 1
cm thick.

Any suggestions would be welcome.

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: morilak-at-cmgm.stanford.edu (David Morilak)
Date: Fri, 22 Oct 1993 12:03:49 PDT
Subject: N-CAM monoclonals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


FYI regarding N-CAM monoclonals:

There are two articles in the most recent J. Comp. Neurol. (336, no. 4), both
of which describe using N-CAM monoclonals in tissue (tongue) that has been
perfusion-fixed with 4% para-formaldehyde:

Smith, Akeson and Shipley, JCN 336(4): 493-506
Nelson & Finger, p.p. 507-516

Hope this helps!

David Morilak
Dept of Psychiatry
Stanford Univ
morilak-at-cmgm.stanford.edu
-------




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sat, 23 Oct 1993 12:23:56 -0500 (CDT)
Subject: Thickness Meassurements

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



RE: Thickness of Surface Layers

John

The obvious answer is making a cross-section of your
specimen and then looking in the appropriate type of microscope
(optical, SEM, TEM.....). What material are you looking at?
What thickness is the surface layer. Can you x-section the samples
or are they unique? Can you use a SIMS system to image the surface
(low res) and then sputter through.....?

Nestor Z.
ANL EMcenter




From: Richard Thornton :      r.thornton-at-trl.oz.au
Date: Mon, 25 Oct 1993 16:18:10 +1000 (EST)
Subject: Infrared Microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

We are currently looking at the purchase of an infrared microscope to aid
us in our failure analysis of integrated circuits. I was wondering if any
one had any experience with these beasts and had any comments to make
about any of the currently available models? Also, what manufacturers
currently make these microscopes, as we are a little in the back waters
here this type of information is a little hard to come by.

Our interest is in microscopy in the wavelength region of 800-1800nm and
as high a resolution as possible.

Thank you for any help



------------------------------------------------------------------------------

Dr. Richard Thornton Telephone: (03) 253 6475
Semiconductor Failure Analysis and Reliability,
Optoelectronics Section, Fax. (03) 253 6664
Telecom Australia Research Labs.,
770 Blackburn Rd., email: r.thornton-at-trl.oz.au
Clayton 3168,
Australia.






From: William Clark :      CLARK-at-kcgl1.eng.ohio-state.edu
Date: Mon, 25 Oct 1993 09:26:42 -0400 (EDT)
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please

unsubscribe William Clark




From: dfa-at-rabbit.physiol.unimelb.edu.au (David Abbott)
Date: Wed, 20 Oct 1993 10:30:33 +1000 (EST)
Subject: Electron microscopy list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reply_ RE} Electron microscopy list
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
Hi everyone?
Why not identify your messages to the list with OM, SEM, TEM, STM, AFM, etc. in
the title? That way we can delete the messages tat are of little or no
interest to us.

--------------------------------------

My personal reason for suggesting this is that, not being
involved in electron microscopy myself, I simply don't have
the time to look through all the EM messages.

Of course, those interested in all topics could subscribe
to both lists and be no worse off than now.

-David.
- -
David Abbott | Email: dfa-at-physiol.unimelb.edu.au
School of Physics & Dept. of Physiology | Phone: +61-3-344-5465
The University of Melbourne, Australia. | Fax: Available on request.


------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;19 Oct 1993 21:25:35 U
Received: by totalrecall.rs.itd.umich.edu (5.67/2.2)
with X.500 id AA15647; Tue, 19 Oct 93 21:23:21 -0400
Received: from anlemc.msd.anl.gov by totalrecall.rs.itd.umich.edu (5.67/2.2)
with SMTP id AA15642; Tue, 19 Oct 93 21:23:19 -0400








From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Mon, 25 Oct 1993 10:47:28 -0600
Subject: Re: Thickness of Surface Layers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} RE: Thickness of Surface Layers
}
} John
}
} The obvious answer is making a cross-section of your
} specimen and then looking in the appropriate type of microscope
} (optical, SEM, TEM.....). What material are you looking at?
} What thickness is the surface layer. Can you x-section the samples
} or are they unique? Can you use a SIMS system to image the surface
} (low res) and then sputter through.....?
}
} Nestor Z.
} ANL EMcenter

Yes, one solution would be to section the material, but that wouldn't allow
mapping of thicknesses over the surface. This work would be for a client
who wants a non-destructive technique. They just want to know the
thickness, and not remove it.

One technique suggested by another lab here at CSU is ellipsometry. As I
understand it, this is a technique that uses polarized light to determine
the thickness of a thin film over a substrate. I don't know what the
mapping or resolution limits are. Unfortunately, I don't know the
approximate thickness, although I think it's 1-100 nm.

It's becoming apparent to me that I need to find out more about these
specimens before I'll be able to help them very much and direct them to the
right technique.

Thanks for all the suggestions.

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 25 Oct 1993 13:07:20 -0500 (CDT)
Subject: LM/counterstaining for immuno/

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



All Microscopy Subscribers:

At the present we only operate one list called Microscopy
and have had one vote/request to split off a portion to an optical
microscopy list with another vote/comment not to do this.
One reason for the split off this is that some users
are only interested in optical or electron or whatever
based microscopy and have no interest in the other fields.
Until I get more feedback we will keep the single list, however,
John Mansfield has recently made a good suggestion relative to
titling your messages to the Microscopy Listserver, which may
improve the operation. Basically he suggested that we add
a label indicating the type of microscopy to the message title.

I like John's Idea. So please from now on when
entering your Subject/Title of your EMAIL Message,
please try to begin your title with an obvious abbreviation
for the type of microscopy you will be addressing.

Here's some suggestions (which of course are not binding
and only reflect my ignorance if the abbreviations
are not current).

LM: Light Microscopy
CFM: Confocal Microscopy
SEM: Scanning Electron Microscopy
TEM: Transmission Electron Microscopy
AEM: Analytical Electron Microscopy
AFM: Atomic Force Microscopy
STM: Scanning Tunneling Microscopy
uAnaly: Microanalysis
Gen: General Questions Comments etc..

Other suggestions? Basically use something obvious that
makes sense.

A typical message title might look something like this

To: microscopy-at-anlemc.msd.anl.gov
/\ /\
|| ||
Abbreviation Subject Matter

Again I have no problems in creating additional lists which
are directed toward a specific microsocpy/microanalysis
methodology. It only requires a sufficient audience.

=======================
Nestor Z.
ANL EMCenter
Email:Zaluzec-at-ANLEMC.MSD.ANL.GOV




From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 25 Oct 1993 11:50:56 -0700 (PDT)
Subject: Re: counterstaining for immunoLM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-carson.u.washington.edu

Try short times in diluted hematoxylin or thionin. Thionin will pick up
more of the cytoplasmic RNS, but we use it routinely, at 4-fold dilution
of the Wisconsin thionin recipe for DAB counterstain and for
autoradiography. Just keep the staining time short. We have also used
methyl green for nickel enhanced DAB and fast green FCF.
If working in B&W, try a green filter or light blue filter in the light
path to enhance the DAB contrast.

Glen MacDonald
Hearing Development Laboratories
University of Washington






From: William Clark :      CLARK-at-kcgl1.eng.ohio-state.edu
Date: Mon, 25 Oct 1993 17:29:56 -0400 (EDT)
Subject: EM microscope list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please

unsubscribe William Clark




From: Gerald Little :      ANGJL-at-medicine.newcastle.edu.au
Date: Tue, 26 Oct 1993 10:25:44 GMT+1000
Subject: Gen: Titles to Microscopy.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.931026102544.352-at-medicine.newcastle.edu.au}

G'day,

In relation to the previous messages concerning the possible
splitting of the group I would strongly suggest that we stay as one
group and definitely adopt the use of appropriate titles to identify
the message as suggested by Nestor Zaluzec. I use TEM, LM and CLSM
and occasionally SEM for my research and having access to the group
allows me to keep up to date with what is going in all these fields.

Regards,
Gerry Little.


Dr Gerald Little | Ph (049) 215618
The Neuroscience Group |
Discipline of Anatomy | Fax (049) 216903
Faculty of Medicine |
The University of Newcastle | Email ANGJL-at-medicine.newcastle.edu.au
Australia, 2308 |




From: Darcy Clark :      03249587-at-mama.minmet.uq.oz.au
Date: Tue, 26 Oct 1993 17:33:04 +1000 (EST)
Subject: TEM: High Resolution TEM of Polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm interested in hearing from anyone who is involved in TEM of polymers,
particularly high resolution. I'm also interested in hearing about any
methods of detecting small crystalline regions in amorphous matrices,
either during or after imaging.
Darcy Clark
03249587-at-minmet.uq.oz.au
Dept.Mining & Metallurgical Eng.
Uni.of QLD. BRISBANE AUSTRALIA.







From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 26 Oct 1993 9:40:47 -0500 (CDT)
Subject: TEM: Polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




We're not doing anything in the way of HREM of polymers here
at ANL, however, I vaguely remember some work done by John Hunt
of Lehigh University on a VG HB501 STEM. It was published I
believe in Ultramicroscopy or the Microbeam Analysis Society
Journal (possibly with Dave Williams) within the last three
years. He was doing combined HR Analytical Microscopy
and imaging.. If I run across the reference I'll post it

Nestor Z.
ANL EM Center




From: :      tonygr-at-EAGLE.MIT.EDU
Date: 26 Oct 93 12:22:18 EDT
Subject: TEM-HREM of polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-------------

Ned Thomas and Weiping Zhang did some very nice HREM imaging of polymers
here at MIT (Ned was also involved with a postdoc at Amherst before he came
to MIT). The people at Akashi (now TOPCON) took some excellent images
of the lattice of Ned's crystalline polymers when I was at their lab trying
out the 002B microscope. I don't know where Weiping is now, but if you
want to write to Ned, his address is:

Prof E. L. Thomas
MIT Room 13-5094
Cambridge MA 02139

If you want to send an E-mail message to him, address it to me and I will
forward it. (I haven't asked his consent to put his E-mail address on this
forum, although I think it is probably public knowledge).

Tony Garratt-Reed






From: Denis Baskin :      baskindg-at-u.washington.edu
Date: Tue, 26 Oct 1993 18:59:55 -0700 (PDT)
Subject: Re: LM cryoimmunofluor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reply_ RE} "Polymer Microscopy," Sawyer & Grubb
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
There is apparently a new edition of this book in production and so those of
you having difficulty getting copies should perhaps wait until the new ediition
hits the bookshelves.

--------------------------------------

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;26 Oct 1993 17:58:46 U
Received: by totalrecall.rs.itd.umich.edu (5.67/2.2)
with X.500 id AA19483; Tue, 26 Oct 93 17:56:44 -0400
Received: from anlemc.msd.anl.gov by totalrecall.rs.itd.umich.edu (5.67/2.2)
with SMTP id AA19478; Tue, 26 Oct 93 17:56:42 -0400
Message-Id: {9310262156.AA19478-at-totalrecall.rs.itd.umich.edu}

We have had the same problem here with cardiac muscle from several
species. The background fluorescence is seemingly impossible to get rid of
in cryo or paraffin sections fixed in 4% paraform. in 0.1 M PB. Any
suggestions?

Denis Baskin
VA Medical Center, Seattle

On Tue, 26 Oct 1993, Steve Barlow wrote:

} Tissue cryosections and immunofluorescence.
} Can anyone suggest some treatments for reducing background
} fluorescence in rabbit muscle tissue? We are using either a 3%
} formaldehyde/PBS fix (3 hours) followed by storage in 0.5% formaldehyde in
} PBS or McClean's sodium periodate/ formaldehyde/ lysine fixative. We are
} using EM grade methanol free formaldehyde from vials. After
} cryosectioning sucrose embedded frozen tissue, we have tried treating with
} 0.2M glycine or 50 mM ammonium chloride for 30-60 min. . After these
} treatments, we still have pronounced tissue fluorescence. I've seen in
} the literature that sodium borohydride is used, but its explosive
} characteristics and its possible alteration of tissue antigenicity seem to
} make it a less desireable treatment.
}
} Any suggestions would be appreciated.
} ----------------------------------------------------------
} Dr. Steven Barlow
} EM Facility/Biology Dept.
} San Diego State University
} San Diego CA 92182-0057
} phone: (619) 594-4523
} fax: (619) 594-5676
} email to sbarlow-at-sunstroke.sdsu.edu
}







From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 28 Oct 1993 09:50:31 U
Subject: Sawyer & Grubb

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Mansfield
North Campus Electron Microbeam Analysis Lab
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or
John.F.Mansfield-at-umich.edu Time: 9:44

Date:10/28/93
NC EMAL

Just my 10cents worth
John Mansfield






From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Thu, 28 Oct 1993 10:08:46 -0500
Subject: Re: LM cryoimmunofluor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




} I've seen in
} the literature that sodium borohydride is used, but its explosive
} characteristics and its possible alteration of tissue antigenicity seem to
} make it a less desireable treatment.
}
I have used sodium borohydride frequently for this purpose and it
has never yet blown up on me. We have also used immunocytochem on anumber
of antigens successfully. So, I don't think its terrrible for antigenicity.
If you are pushing things, there may be a small detrimental effect. I use a
1 mg/ml solution in PBS. You have to make up the solution just before use.
We keep the powder in a dessicant at room temp.

In one lab I was in, they used to cut the PBS 1:1 with methanol. This
slowed down the reaction. But, methanol itself can do some nasty things to
your tissue. A ten to 30 min. incubation is typical.Some folks use a few
changes. Ie, incubate for 10 min and then change to fresh solution for
another 10.

However, my understanding was that NaBH4 works for reducing
autofluor from unreacted aldehydes in gluteraldehyde fixation. That is,
glut is bi functional, so there can be dangling aldehydes out there. These
the borohydride attacks and prevents from forming highly conjugated
somethings or other, which are autofluor. In my work, on plant tissue, no
help from NABH on autofluor when only pfa is used. Of course, this is no
guarantee your tissue's autofluor might not be eased with the NABH4.

Hope this helps,
Tobias Baskin
******************************** ***************
Tobias I. Baskin /~~~\
Biol. Sci's * Univ. of Missouri c|o o\
Columbia, MO 65211 USA \ = /
Tel:314-882-0173 """
FAX 314 - 882 - 0123
baskin-at-biosci.mbp.missouri.edu





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Thu, 28 Oct 1993 19:58:22 -0500 (CDT)
Subject: LM-cryoimmunofluor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Steve Barlow asked
} Tissue cryosections and immunofluorescence.
} Can anyone suggest some treatments for reducing background
} fluorescence in rabbit muscle tissue?.........(text deleted)


I have absolutely no experience in LM Fluorescence problems
but an obvious solution to me (which probably means I'm wrong)
is to change the wavelength of the light source. If the illumination
system is causing the staining(?) chemical to fluoresce then
might not it be possible to change the excitation source? Presumably
the staining chemical is sensitive to the radiation you are using
thus rather than move to a potential dangerous chemical try
changing the source.....

Now somebody please cure my ignornace and
tell me why this is probably wrong....

Nestor Z.
ANL EM Center




From: :      morilak-at-cmgm.stanford.edu
Date: Thu, 28 Oct 1993 10:08:13 PDT
Subject: Re: LM cryoimmunofluor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Steve:

Regarding the use of sodium borohydride, I have used it regularly as a
pretreatment for immunoperoxidase when using glutaraldehyde in my fix - it's
great at reducing background (presumably by reducing double bonds - potential
sites of non-specific DAB oxidation), and also seems to unmask antigenicity. I
have been able to increase my primary antibody dilutions several fold in some
cases. I have not noticed any difference though when using only
paraformaldehyde, so I'm not sure it would help with your background
fluorescence problem.

After initial PBS washes

Incubate 30 min in 0.5% NaBH4 in 50 mM Tris, pH 7.4

Then 3 x PBS washes before blocking serum

The borohydride solution should be weighed out and put into solution
immediately before use (it goes in fast). It does generate a lot of bubbles,
so use plenty of volume on your tissue.

David Morilak
Dept of Psychiatry
Stanford University
morilak-at-cmgm.stanford.edu
-------




From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Fri, 29 Oct 1993 11:11:31 -0400 (EDT)
Subject: Re: LM cryoimmunofluor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} } I've seen in
} } the literature that sodium borohydride is used, but its explosive
} } characteristics and its possible alteration of tissue antigenicity seem to
} } make it a less desireable treatment.
} I have used sodium borohydride frequently for this purpose and it
} has never yet blown up on me. We have also used immunocytochem on anumber

Just a warning: green Bodipy does not work after this treatment.






From: MN :      morilak-at-cmgm.stanford.edu
Date: Fri, 29 Oct 1993 10:58:42 PDT
Subject: Re: LM-cryoimmunofluor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On 10-29, Nestor Zaluzec writes:

I have absolutely no experience in LM Fluorescence problems
but an obvious solution to me (which probably means I'm wrong)
is to change the wavelength of the light source. If the illumination
system is causing the staining(?) chemical to fluoresce then
might not it be possible to change the excitation source? Presumably
the staining chemical is sensitive to the radiation you are using
thus rather than move to a potential dangerous chemical try
changing the source.....

Nestor:

I'm not sure if this is the answer you're looking for, but here's my shot
anyway: The application, if I understand Steve correctly, is in using a
fluorescent-conjugate secondary antibody to detect a primary antibody in
immunohistochemistry. The conjugates are typically fluorescein (FITC) and
rhodamine (TRITC, a fluorescein derivative). The light source is a high energy
mercury lamp, and the light passes through a filter cube that is matched to
the fluor that is being examined. These cubes are commercially available - the
best are probably from Nikon in my experience. The excitation, emission and
notch filter combinations result in a fairly tight wavelength window, which is
necessary to be able to discriminate the different tags with different filters
(for co-localization studies) without bleed-through, since their spectra do
overlap considerably. The problem is that, using the excitation/emission
settings necessary for viewing FITC and TRITC, there may be some tissue
elements that are also excited, or that emit autofluorescence. (A note - never
use Kim-Wipes on your slides - they are treated with something, presumably to
make them nice and white, that demonstrates a most impressive
"autofluorescence"!). So, the options are to switch fluorescent labels to get
away from the wavelengths that cause fluorescence in tissue (but the alternate
choices and reagents are extremely few), or reduce the tissue fluorescence by
a variety of pretreatments.

Here's another suggestion: I have used a mounting medium made from
"antifadent" tablets from CitiFluor (City University, London). The mountant is
essentially 90% glycerol/PBS with 1 dissolved tablet/10 ml. The tablet is
presumably some proprietary detergent concoction, but it does help reduce
background fluorescence. It's fairly expensive though - as I remember about
$100 (US) for 10 tablets, so I only use it for "critical" work (but isn't it
all?).

I hope this is informative!

David Morilak
Department of Psychiatry
Stanford University
morilak-at-cmgm.stanford.edu
-------




From: Denis Baskin :      baskindg-at-u.washington.edu
Date: Sat, 30 Oct 1993 13:37:05 -0700 (PDT)
Subject: Re: LM: Immuno controls

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: baskindg-at-carson.u.washington.edu

Steve:
The rule of thumb that we use for polyclonal abys when preimmune serum
from the same rabbit is not available is to use normal rabbit serum at the
same diluton as the primary aby. However, you may have to screen several
batches of normal rabbit serum to find one that has low nonspecific
binding. Monoclonal aby concentratons are usually given in mg/ml or
something like that, and you usually dilute them for use. If, for
example, you use a mca at 0.01 mg/ml, then the control would be mouse IgG
at the same concentration, although you can also dilute normal mouse serum
to the same concentration. The problem we frequently encounter is that
normal mouse serum tend to be "sticky" and you may have to screen a dozen
samples to find one without a lot of nospecific staining. Other labs have
found similar problems with NMS. Just replacing the primary aby with PBS
is definitely not adequate!. If you have to use the primary abys at c.a.
1:100 or 1:50 or less, you may be introuble since the NSB can be as high
as the SB. All the more reason to serach for a good normal serum with low NSB.
Good luck.
Denis Baskin
Seattle VA Medical Center
206-764-2138

On Thu, 28 Oct 1993, Steve Barlow wrote:

} 1. what does one use for a control when using a monoclonal Ab--there is,
} in this case, no 'pre-immune'? Is a monoclonal Ab known not to be expresed
} in the sample the best alternative?
}
} 2. what does one use for a control when using a polyclonal rabbit Ab
} (commercial or otherwise) for which no pre-immune is available?
} My user is not satisfied using PBS alone in place of a primary. If I use
} rabbit normal serum or rabbit IgG--how does one come up with a meaningful
} concentration? Presumably this control is to show how 'sticky' the sample
}