From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 1 Nov 1993 17:17:45 -0800 (PST)
Subject: Re: LM-cryoimmunofluor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



On Thu, 28 Oct 1993, Nestor J. Zaluzec (708)-252-5075, -4964 wrote:

}
} } Steve Barlow asked
} } Tissue cryosections and immunofluorescence.
} } Can anyone suggest some treatments for reducing background
} } fluorescence in rabbit muscle tissue?.........(text deleted)
}
}
} I have absolutely no experience in LM Fluorescence problems
} but an obvious solution to me (which probably means I'm wrong)
} is to change the wavelength of the light source. If the illumination


I tried to avoid autofluorescence by changing the wavelength. I suppose
that might work in some situations where a single specific component with
a narrow band of excitation or emission is present. But some things, like
the lipofuscin granules in gut will fluoresce at any wavelength.
my hypothesis is that most biological autofluoresce is due to a number of
molecules (and processing distortions) that will glow over a range of
frequencies.

For what its worth, my recent experience with whole vestibular organs
showed that FITC, AMCA and Cascade Blue wash out quickly, leaving little
backgound. But Texas Red required an overnight wash.






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 1 Nov 1993 17:00:24 -0800 (PST)
Subject: Re: LM: Immuno controls

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-carson.u.washington.edu

} Some general questions for discussion:
}
} 1. what does one use for a control when using a monoclonal Ab--there is,
} in this case, no 'pre-immune'? Is a monoclonal Ab known not to be expresed
} in the sample the best alternative?
}
} 2. what does one use for a control when using a polyclonal rabbit Ab
} (commercial or otherwise) for which no pre-immune is available?
My 2 cents:
I had gotten by with finding the
IgG concentration of the antibody preparation. Then I dilute a normal
(non-immune) serum, or purified IgG, from the appropriate species to a
similar concentration. IgG and IgM estimates for the normal sera of several
species may be found in R. Lindmark, et al, J. Immunmol. Meth., 62:1-13,
1983. Of course, not all vendors know that much about their products, and
most homemade antibodies are relatively uncharacterized. Sometimes the
IgG content of an immune serum or monoclonal prep. can be estimated by
comparing the total protein to that of non-immune serum for that
species, or control culture/ascites fluids.
I have encountered commercial "non-immune" sera that gave a better
anti-cytokeratin label than the monoclonals being evaluated.








From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 2 Nov 1993 11:01:10 -0600 (CST)
Subject: LM: What is IgG....

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I've seen alot of discussion about staining lately which
is interesting, but now I have a simple question. We deal mainly
with metals/ceramics/semiconductoring materials hence, we do no staining etc..
So for the "novice in me" can someone tell me what IgG and IgM are?
I'm not able to deduce the meaning from the context of the discussion.
It would help me (and probably a few others) to follow the discussion
a little better.

Remember this list has a whole slew of different readers and not all
of us know each others "standard" abbreviations from the different
fields of microscopy (or even the same field but in different applications).
Try to keep this in mind when posting things.

Nestor Z.
ANL EM Center




From: '( :      morilak-at-cmgm.stanford.edu
Date: Tue, 2 Nov 1993 11:50:52 PST
Subject: Re: IgG

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nestor:

IgG and IgM stand for immunoglubulin G and immunoglobulin M, and in this
context refer to different types of antibodies used for immunohistochemistry.
For most applications, primary antisera are of the IgG type, but for some
applications, and for antibodies derived from certain sources and by different
procedures, they may be IgM. This is important to know, since the secondary
antibody, which is directed against the primary antibody and usually carries
the tag which will allow visualization, must be directed against the
appropriate immunoglobulin and the right host species.

There is an excellent review, both for its information content and its
historical value, on the basics of indirect immunohistochemiostry by Ludwig
Sternberger, 1974. I don't have the reference handy, but I will track it down
and report back.

Hope this has been useful...

David Morilak
Dept of Psychiatry
Stanford university
morilak-at-cmgm.stanford.edu
-------




From: ST :      morilak-at-cmgm.stanford.edu
Date: Tue, 2 Nov 1993 11:54:28 PST
Subject: Re: LM- what is IgG...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nestor:

IgG and IgM stand for immunoglubulin G and immunoglobulin M, and in this
context refer to different types of antibodies used for immunohistochemistry.
For most applications, primary antisera are of the IgG type, but for some
applications, and for antibodies derived from certain sources and by different
procedures, they may be IgM. This is important to know, since the secondary
antibody, which is directed against the primary antibody and usually carries
the tag which will allow visualization, must be directed against the
appropriate immunoglobulin and the right host species.

There is an excellent review, both for its information content and its
historical value, on the basics of indirect immunohistochemiostry by Ludwig
Sternberger, 1974. I don't have the reference handy, but I will track it down
and report back.

Hope this has been useful...

David Morilak
Dept of Psychiatry
Stanford university
morilak-at-cmgm.stanford.edu
-------




From: morilak-at-cmgm.stanford.edu (David Morilak)
Date: Tue, 2 Nov 1993 12:18:37 PST
Subject: Re: LM - What is IgG....

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I've had some real problems trying to send this message to the list, so I hope
it gets through this time, and please forgive me if it gets sent a few times:

Nestor:

IgG and IgM stand for immunoglubulin G and immunoglobulin M, and in this
context refer to different types of antibodies used for immunohistochemistry.
For most applications, primary antisera are of the IgG type, but for some
applications, and for antibodies derived from certain sources and by different
procedures, they may be IgM. This is important to know, since the secondary
antibody, which is directed against the primary antibody and usually carries
the tag which will allow visualization, must be directed against the
appropriate immunoglobulin and the right host species.

There is an excellent review, both for its information content and its
historical value, on the basics of indirect immunohistochemiostry by Ludwig
Sternberger, 1974. I don't have the reference handy, but I will track it down
and report back.

Hope this has been useful...

David Morilak
Dept of Psychiatry
Stanford university
morilak-at-cmgm.stanford.edu
-------




From: morilak-at-cmgm.stanford.edu (David Morilak)
Date: Tue, 2 Nov 1993 17:10:52 PST
Subject: Re: LM- What is IgG

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0ouTk2-000dBZC-at-unpsun1.cc.unp.ac.za}
To: microscopy-at-ANLEMC.MSD.ANL.GOV

Following up my earlier reply, here is the citation for Sternbergers classic
reference on immunocytochemical techniques:


Sternberger, Ludwig A. IMMUNOCYTOCHEMISTRY (Englewood Cliffs, N.J. :
Prentice-Hall, 1974)

Sternberger, Ludwig A. IMMUNOCYTOCHEMISTRY. 2nd ed. (New York : Wiley,
1979)

Sternberger, Ludwig A. IMMUNOCYTOCHEMISTRY. 3rd ed. (New York : Wiley,
1986)

If anyone is interested, it's a great introduction to immunocytochemistry, and
for anyone who is doing it, it is a great look into how the technique has been
developed.

David Morilak
morilak-at-cmgm.stanford.edu

-------




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 2 Nov 1993 20:55:38 -0600 (CST)
Subject: LM:IgG & Cruise Missles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Thanks to all for the general info on IgG & IgM.
Just goes to show you that us materials
types are interested in lots of things too!

BTW I smiled at Michael Herron's closing comment:

} Hey, it ain't no cruise missle but it what we do :^)

Remind me in a less busy time to describe how the ANL EM
Center proposed and developed a protocol for using an SEM on Cruise
and other Missles (US & Russian) as a nondestructive
means for nuclear arms verification. It was actually a small ($20K)
but funded program for a year by the START program! The funny thing
is the procedure worked. It got killed in the end because
anything that inexpensive couldn't work (at least
one of the explaination I got)

:-)

Nestor Z.
ANL EM Center





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 3 Nov 1993 16:25:09 -0600 (CST)
Subject: ALL Microscopy: Image Storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0ouTLM-000dBTC-at-unpsun1.cc.unp.ac.za}
To: microscopy-at-ANLEMC.MSD.ANL.GOV




On Wed, 3 Nov 1993, Michael Cammer wrote:

}
} We are looking for a system for gross storage and easy retrieval of a
} large number of images for both teaching and research. Does anybody have
} any suggestions?
}
} Basically, we have a large (microscope) slide library that we would like
} to scan in color. We need the images available for playback on a video
} monitor and for photography. Should we use a computer and, if so, with

} what software (e.g. database) and what storage device (for speed, quanity,
} compatibility w/ other systems, and an eye towards support in the future)?
} The initial application is for a slide library for teaching, but, perhaps,
} this could be turned into an interactive system available to students.
} Also, I have been thinking that such a system might be useful for storing
} our images for research; even without this specific query, we need more
} and more compatible storage.
}
} Any suggestions for specific parts or for whole integrated systems (e.g.
} is there something smilar for other fields?) would be welcome.
}
} Thanks-
}
} Michael cammer-at-aecom.yu.edu
}
}
}

Cost aside, one alternative is HDTV for capturing the images. I'm not
sure what the current state of the art is, but several years ago, Sony
offered complete HDTV systems (cameras, analog and digital storage,
post-processing hardware, optical fiber closed-circuit transmission, etc.)
specifically aimed at the pathology market. I saw the system
demonstrated in the Sony showroom in Tokyo; the image quality was
extraordinary. As I recall, there are several sites in the US that
offer HDTV-based remote consulting.

Apart from cost, one reason to hold off on the Sony system is that their
HDTV standard is different than the one that three US developers have
decided (at long last) to develop jointly: You could end up with a pricey
white elephant.

Storage of massive amounts of data will depend on a number of factors
that include required speed of retrieval, number of users who need
simultaneous access, available hardware, money. Running all the images
onto CD-ROM is not a bad alternative; costs are not terribly bad (if
you want to make your own, you'll have too pay around $4K) although speed
of retrieval may be frustrating.

I'm quite interested in hearing of ways people have developed archives of
microscopy data. Having representative images on-line in conjunction with
other data can make a very powerful tool.

Dave Coder
Dept. of Immunology
Internet: dcoder-at-u.washington.edu






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Wed, 3 Nov 1993 15:01:40 -0800 (PST)
Subject: Re: ALL Microscopy: Image catalogs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-carson.u.washington.edu

Related to the current discussion thread on imarge archiving - what
systems and software do people use for cataloging and retrieving images?
We are accumulating stacks of images from calcium ratioing, confocal,
serial images from convential microscopy (lm and SEM) and stuff that has
been converted to projection slides and Linotronic files.

There are both commercial and shareware programs offered.

Glen MacDonald
Hearing Development Laboratories RL-30
University of Washington
Seattle, WA 98195
(206) 543-8360





From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 3 Nov 1993 18:29:55 U
Subject: Critical Point Dryer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Mansfield
North Campus Electron Microbeam Analysis Lab
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or
John.F.Mansfield-at-umich.edu Time: 17:27

Date:11/3/93
NC EMAL






From: Jouko K. M„ki :      jokamaki-at-utu.fi
Date: Thu, 4 Nov 1993 08:48:43 +0200
Subject: Re: Critical Point Dryer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Warning: Original message contained 8-bit characters, however during
the SMTP transport session the receiving system was unable to
announce capability of receiving 8-bit SMTP (RFC 1425-1428),
and as this message does not have MIME headers (RFC 1341) to
enable encoding change, only option was to STRIP sent characters
to 7-bit :-(


On Thu, 4 Nov 1993 04:29:55 +0200, John Mansfield wrote:

} Date:11/3/93
} NC EMAL
} Subject: Critical Point Dryer
} I have a colleague who is in the market for a critical point drying system.
} Not knowing anything about such things or exactly how they work, I want to ask
} if there is a manufacturer or model that is preferred by the majority of people
} and if it is possible to dry material that is wet with a liquid other than
} water.
} Any help or info would be appreciated.
} Thanks.
} John Mansfield

The idea of critical point drying is to change the liquid to a suitable
liquified gas, which has a low enough critical point to change directly
fron liquid phase to gas phase. It does not matter wether the change is
made directly or via a series of liquids. The only thing which matters is
to maintain the sample as intact as possible while changing the liquids. So
all you need to know is the solubility of different liquids in each other
and finally in the liquified gas.

Hopefully this helps, others can give you more exact advice.

Best regards,

Jouko Mki

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Jouko K. Mki Navigare necesse est...
Laboratory Manager, Ph.D.
Laboratory of Electron Microscopy Kiinamyllynkatu 10 FIN-20520 TURKU FINLAND
University of Turku Tel.: + 358 21 633 7318
INTERNET: jouko.maki-at-utu.fi FAX.: + 358 21 633 7380




From: Jouko K. Maki :      jokamaki-at-utu.fi
Date: Thu, 4 Nov 1993 10:08:44 +0200
Subject: Re: Critical point drying

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Thu, 4 Nov 1993 04:29:55 +0200, John Mansfield wrote:

} Date:11/3/93
} NC EMAL
} Subject: Critical Point Dryer
} I have a colleague who is in the market for a critical point drying system.
} Not knowing anything about such things or exactly how they work, I want to ask
} if there is a manufacturer or model that is preferred by the majority of people
} and if it is possible to dry material that is wet with a liquid other than
} water.
} Any help or info would be appreciated.
} Thanks.
} John Mansfield

The idea of critical point drying is to change the liquid to a suitable
liquified gas, which has a low enough critical point to change directly
fron liquid phase to gas phase. It does not matter wether the change is
made directly or via a series of liquids. The only thing which matters is
to maintain the sample as intact as possible while changing the liquids. So
all you need to know is the solubility of different liquids in each other
and finally in the liquified gas.

Hopefully this helps, others can give you more exact advice.

Best regards,

Jouko Maki

P.S. Sorry for repeating this message. The first sending contained 8-bit
characters causing errors. I hope this goes through properly.
jm

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Jouko K. Maki Navigare necesse est...
Laboratory Manager, Ph.D.
Laboratory of Electron Microscopy Kiinamyllynkatu 10 FIN-20520 TURKU FINLAND
University of Turku Tel.: + 358 21 633 7318
INTERNET: jouko.maki-at-utu.fi FAX.: + 358 21 633 7380




From: NoelGoldsmith:      mt1ntg%fencer.cis.dsto.gov.au-at-relay.tc.umn.edu (Noel
Date: Wed, 3 Nov 1993 19:40:03 -0600
Subject: Re: system for storing lots of images?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Received: From ANLVM(MAILER) by ANLEMC(ANJE5.40) for ZALUZEC-at-ANLEMC; Wed, 3 Nov
93 20:42 CST
Received: from ANLVM by ANLVM (Mailer R2.07B) with BSMTP id 6784; Wed, 03 Nov
93 19:43:46 CST
Received: from nx1.soils.umn.edu by ANLVM.CTD.ANL.GOV (IBM VM SMTP R1.2.2ANL-MX)
with TCP; Wed, 03 Nov 93 19:43:44 CST
Received: from (localhost) by nx1.soils.umn.edu (5.65c)
id AA12448; Wed, 3 Nov 1993 19:40:03 -0600
Goldsmith)
To: Multiple recipients of list {nih-image-at-nx1.soils.umn.edu}

} We are looking for a system for gross storage and easy retrieval of a
} large number of images for both teaching and research. Does anybody have
} any suggestions?
}
} Basically, we have a large (microscope) slide library that we would like
} to scan in color. We need the images available for playback on a video
} monitor and for photography. Should we use a computer and, if so, with
} what software (e.g. database) and what storage device (for speed, quanity,
} compatibility w/ other systems, and an eye towards support in the future)?
} The initial application is for a slide library for teaching, but, perhaps,
} this could be turned into an interactive system available to students.
} Also, I have been thinking that such a system might be useful for storing
} our images for research; even without this specific query, we need more
} and more compatible storage.
}
} Any suggestions for specific parts or for whole integrated systems (e.g.
} is there something smilar for other fields?) would be welcome.
}
} Thanks-
}
} Michael cammer-at-aecom.yu.edu
Michael,
We are also looking at alternatives for the long term archiving and easy
retrieval of digital images.
Many of our images are directly aquired using Perceptics HR 24 RGB frame
grabber with a SONY DXC930P (P for Pal) which gives about 1Mb per image of
real 24 bit colour with low noise if the lighting is good. Or we use a
Videk (Kodak) MegaVision with a Perceptics MegaGrabber for a 1.4Mb
Monochrome image, or we use the Data Translation DT 2555 Quik Capture for a
400Kmonochrome image. These images are used for illustration, measurement
etc. We have a Richoh read-write optical disc which is useful for storage,
and works well. We have now started to worry about back-ups of our now
extensive files of images. We are going to make a Photo CD, which will
require the production of films or slides from digital images. although
next year Kodak will offer a service called portfolio which will allow the
production of a CD containing digital only material.
The idea of a photo cd is attractive because it is becoming a well accepted
plateform for images, and is likely to be durable in the physical and
temporal sense. The other advantages are the availability of "juke boxes"
for CD's allowing the use of six or so in one machine. This access might
not be as quick as from a hard disc, but it will be better than scanning in
the original again.
Because photo-cd has the "thumbnails", rapid data base access becomes a
reality, software is already available, eg shoebox. So we will report tto
the list as we learn.
Noel T Goldsmith
DSTO
Aeronautical Research Laboratory
506 Lorimer Street
Port Melbourne
Vic 3207
Australia





From: John Ladwig :      jladwig-at-soils.umn.edu
Date: Wed, 3 Nov 1993 18:51:40 -0600
Subject: Re: system for storing lots of images?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Received: From ANLVM(MAILER) by ANLEMC(ANJE5.40) for ZALUZEC-at-ANLEMC; Wed, 3 Nov
93 19:54 CST
Received: from ANLVM by ANLVM (Mailer R2.07B) with BSMTP id 6649; Wed, 03 Nov
93 18:55:24 CST
Received: from nx1.soils.umn.edu by ANLVM.CTD.ANL.GOV (IBM VM SMTP R1.2.2ANL-MX)
with TCP; Wed, 03 Nov 93 18:55:22 CST
Received: from (localhost) by nx1.soils.umn.edu (5.65c)
id AA12018; Wed, 3 Nov 1993 18:51:40 -0600

} } } } } On Wed, 3 Nov 1993 18:25:54 -0600, kartenh-at-Sdsc.Edu said:

} } Cost aside, one alternative is HDTV for capturing the images.

kartenh} Another alternative is the Kodak digital camera back on a
kartenh} Nikon 8008. This directly dumps data into a Mac through
kartenh} the SCSI port. It provides a color (24 bit?) image with
kartenh} about 1500 pixel horizontal resolution X about 1000
kartenh} pixel. It grabs the image in about the same time that
kartenh} film does. But it is slow in dumping it into the
kartenh} computer. (Each image is about 4.5 MB). They provide a
kartenh} plug-in driver for Adobe Photoshop, so you can directly
kartenh} manipulate the image. H. Karten

I believe this is the very same DCS2000 I used earlier this year. It
is most emphatically *not* a 24 bit scanner, although all the product
lit seems to imply it very strongly. Relevant fragments from old
sci.image-processing articles follow. As I wrote before, the PhotoCD
images I just got were *much* preferable to the DCS2000 images.

|} From: sasrer-at-unx.sas.com (Rodney Radford)
|} Subject: Re: Digital Cameras - Suggestions?
|} Date: Wed, 9 Jun 1993 23:27:46 GMT

|} John Peterson {jp-at-taligent.com} writes:
|} } Kodak DCS. This is a Nikon 35MM with a new back containing a
|} } 1024x1024x24 CCD array. Old models required you to lug around a
|} } suitcase, newer ones have a box about the size of the camera body
|} } attached to the bottom of the Nikon. Suitcase stores about 1-200
|} } photos (depends on compression), smaller one can store photos in
|} } small strap on SCSI disk. Mondo expensive, $8-20K. Smaller
|} } cameras only take 20-30 photos on a battery charge; suitcase
|} } models go longer. Suitcase has an LCD screen to review images.
|}
|} Most of your description was very accurate, and very
|} informative. However, I don't know of any 24bit DCS cameras
|} available (someone correct me if I am wrong, and give me the model
|} number). We currently have a DCS200 camera, which is a Nikon F3
|} camera with a strap on hard drive capable of storing 50 images.
|} However, it is only an 8bit CCD array with an RGB color mask in
|} front so that each pixel can be EITHER red, green, or blue, such
|} as:
|}
|} R G R G R G R G ....
|} G B G B G B G B ....
|} R G R G R G R G ....
|} G B G B G B G B ....
|}
|} The resulting image on the DCS200 is a 1024x1536 image and takes
|} 1.5MB (data is NOT compressed in the camera/disk). After this image
|} is downloaded into the computer, it is interpolated into a full
|} 24bit image by calculating 'suitable' values of red, green, and
|} blue for each pixel. The result is pretty good, but not as good as
|} a real 24bit image!



|} From: sasrer-at-unx.sas.com (Rodney Radford)
|} Subject: Re: Digital Cameras - Suggestions?
|} Date: Thu, 10 Jun 1993 13:05:40 GMT
|}
|} ledwards-at-leland.Stanford.EDU (Laurence James Edwards) writes:
|}
|} } Not to quibble, but in standard usage (at least on the output
|} } side) a 24bit image means 8bits in rgb ... each "pixel" is
|} } composed of an rgb triad. So I would say John had the depth right
|} } but the resolution wrong ... that is, the unprocessed image is
|} } really 512x768x24. Do they use linear interpolation, or something
|} } a little more sophisticated?

|} Except that the interpolation algorithm yields a 1024x1536x24 bit
|} image, not the 512x768x24 image you mention above. So they 'fake'
|} the other 16bits in EVERY pixel location. So, I still stand by my
|} original 1024x1536x8 bit description of the CCD array. Below is a
|} paragraph from the DCS200 Programmer's Specifications:

|} "The CCD sensor for the DCS200 Digital Camera does not capture all
|} three RGB colors for each pixel. Instead, it captures one color for
|} each pixel, so that interpolation is required to construct three
|} full RGB planes for the image. The grid below represents the pixel
|} grid before interpolation. Green pixels predominate the grid because
|} Green captures the luminance levels which can translate across to the
|} Red and Blue planes."

|} The interpolation algorithm consists of three seperate passes over
|} the image, interpolating the green, red and blue values. The Green
|} interpolation algorithm basically looks at the 4 nearest neighbors
|} as two pairs (up/down, left/right) and calculates the two delta
|} values. It then uses the average of the pair with the largest delta
|} value. The red/blue algorithms are very similar, except that there
|} is no test for largest delta values - instead some pixels always
|} use the up/down pair, some use the left/right pair, and others use
|} a 4way average all all neighbors.

|} There is also a black-level color correction pass (enhances
|} contrast), and a gamma correction pass. There is also a list of
|} 'defective' CCD rows in the camera, and how to adjust for them
|} (replace with a neighbor row, or average with two other rows).

|} We have received 3 seperate copies of the KCS200 manual, and in
|} EVERY one, the algorithms have changed, so I guess they are not
|} fixed in stone, and Kodak is still experimenting. Note that it is
|} also possible to just use the interpolated Green plane as a B&W
|} image.




From: rzimm%greenwood.cr.usgs.gov-at-relay.tc.umn.edu (Robert Zimmermann)
Date: Wed, 3 Nov 1993 21:48:58 -0600
Subject: Re: system for storing lots of images?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Received: From ANLVM(MAILER) by ANLEMC(ANJE5.40) for ZALUZEC-at-ANLEMC; Wed, 3 Nov
93 22:51 CST
Received: from ANLVM by ANLVM (Mailer R2.07B) with BSMTP id 7009; Wed, 03 Nov
93 21:52:44 CST
Received: from nx1.soils.umn.edu by ANLVM.CTD.ANL.GOV (IBM VM SMTP R1.2.2ANL-MX)
with TCP; Wed, 03 Nov 93 21:52:42 CST
Received: from (localhost) by nx1.soils.umn.edu (5.65c)
id AA13107; Wed, 3 Nov 1993 21:48:58 -0600

} We are looking for a system for gross storage and easy retrieval of a
} large number of images for both teaching and research. Does anybody have
} any suggestions?
............
} Michael cammer-at-aecom.yu.edu

to which Harvey Karten replied:

} There are several options. Low to medium resolution, use Video. File size
} is modest (circa 300KB per image). Resolution is low = despite what your
} television salesman tells you, you won't do better than about 350 lines of
} resolution. You can then archive them for database using Aldus Fetch or
} Multi-User's Search 2.0 image database.
} Alternately, use intermediate photos, and send them off to Kodak for
} Photo-CD ROMs. They provide a few different levels of resolution (from
} video levels up to about 2,000 lines), depending on your display
} capability. Increasing amounts of software to access Kodak Photo-CDs.
}
} For the user demanding higher levels of resolution, you will have to go to
} a Scitex or Leaf scanner (costs about $15,000) which will give youj about
} 5,000x4,000 pixel resolution. This is slightly better than a sharply
} focussed image on Kodachrome 35 mm (also about 20 megapixels, or 4K x 5K.)
} }
Image database programs for the Mac were reviewed in the Sept. '93 MacUser,
p190, a good starting point for comparisons. "Aldus Fetch" (mentioned by
Harvey) received a four-mice review. Our graphics group here has set up a
very usable multiuser database using Fetch, for drafted figures and scanned
images. It is used to organize and access their archive and active files.

But, how to acquire the digital files? Capture by CCD camera directly from
the microscope will be a problem regardless of it's CCD array size if you
want to capture a large area (ie. low magnification). If you shoot film at
low magnification, you can then scan the transparencies with a film scanner
(2000-4000dpi,big files !), your own (your time and $$) or Kodak's PhotoCD
($$). You do have control of cropping, resolution, color or exposure
control if you do it yourself. If your images have a narrow color range,
you can reduce the file size by x3 by transforming it to 8-bit indexed
color without much perceptible loss. "Photoshop" or "Fast Eddie" can do
this for you. I have had some discussions with people in our photo
library (chemical), who want to establish a digital equivalent. They have
over 50,000 negatives and slides in the freezers. Let Kodak handle it!
(As per Harvey's recommendation). PhotoCD's provide enough resolution
(2000x3000 pixels) for direct 300dpi dye-sublimation output of 6.67"x10"
prints. Discussion of the adequacy of dye-sub output can be found in the
earlier record of this forum. A review of Kodak's image manipulation
program, "Kodak PhotoEdge", can also be found in Sept'93 MacUser. Kodak
PhotoCD's can even be viewed on the home TV, a nice feature for cramming
students, even if not high-res.

I have been playing with a Nikon LS-3510AF film scanner (on loan from my
local Nikon representative, S & M Microscopes, Colorado Springs, CO --A
"thank you" plug). You know what? You can stick a slide (40mm x40mm scan
area) directly in the 35mm slide holder and scan it. No microscope! At
the maximum scanner resolution of 3165 dpi that is a resolution of 0.124
micrometers/pixel which is about what I get at x100 on my microscope
set-up. It does a nice job with a slide sandwiched between crossed
polarizers (we geologists do this a lot). I assume your slides are however
of animal or vegetable, not mineral subjects. You wouldn't be able to do
fluorescence stains without major, but not impossible, modifications to the
scanner ( Is there a market out there?). Will microscopes be obsolete
soon? Within the last year there was an article in "Advanced Imaging"
about the development of a digital microscope. I can not remember the
exact issue off-hand. In that set up the sample is stepped across a high
density (4000-6000 pixel) linear array CCD, just like most film scanners do
( but theirs still looked like a microscope).

Good luck with your project. Maybe some of this will help.



=============================================================
Robert Zimmermann rzimm-at-usgs.gov (Internet)
U.S. Geological Survey 303-236-5626
MS 905 Box 25046 Federal Center
Denver CO 80225





From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Thu, 4 Nov 1993 09:24:23 -0700
Subject: Re: Critical point drying

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} On Thu, 4 Nov 1993 04:29:55 +0200, John Mansfield wrote:
}
} } Date:11/3/93
} } NC EMAL
} } Subject: Critical Point Dryer
} } I have a colleague who is in the market for a critical point drying system.
} } Not knowing anything about such things or exactly how they work, I want to ask
} } if there is a manufacturer or model that is preferred by the majority of
} } people
} } and if it is possible to dry material that is wet with a liquid other than
} } water.
} } Any help or info would be appreciated.
} } Thanks.
} } John Mansfield
}
} The idea of critical point drying is to change the liquid to a suitable
} liquified gas, which has a low enough critical point to change directly
} fron liquid phase to gas phase. It does not matter wether the change is
} made directly or via a series of liquids. The only thing which matters is
} to maintain the sample as intact as possible while changing the liquids. So
} all you need to know is the solubility of different liquids in each other
} and finally in the liquified gas.
}
} Hopefully this helps, others can give you more exact advice.
}
} Best regards,
}
} Jouko Maki

The most common transition fluid used these days is liquid carbon dioxide.
Specimens are typically dehydrated through a graded series of organic
solvent, ethanol or acetone, then placed in the cooled critical point dryer
(CPD). Ethanol and acetone are both miscible with CO2. Once the specimens
are in the chamber, liquid CO2 is introduced, and, through several
exchanges (filling and draining with CO2 under pressure from the tank), the
organic solvent is completely replaced by CO2.

The chamber is kept cool to keep the CO2 in liquid phase. After exchange
of liquids, the temperature is increased and the pressure inside the vessel
increases also. Once the critical point is passed, 1073 psi and 31.1 deg C
for CO2, the vessel is slowly vented. After atmospheric pressure is
reached, the chamber is opened and the specimens removed. They are warm
and dry, and they tend to rehydrate pretty fast, so having a dessicator
handy is a good idea.

The typical CPD device is a high pressure bomb that has valves (3) for
introducing liquid CO2 and venting liquid (bottom of chamber) and gas (top
of chamber). There is a mechanism for heating and cooling, too. One
temperature control system has electronic heating and cooling devices
(can't think of the name now) and sensors for temp control. These can be
automated. The other kind, which I prefer, has a water jacket around the
pressure vessel. You run ice water in to cool and warm/hot water to heat.
You have to keep an eye on the chamber to make sure you don't heat too
fast. But I watch the chamber, even when it's automated.

I like the water controled units because (1) they are cheaper (2) there are
fewer parts to go bad (3) they are smaller. The one I use now is made by
Polaron (model 3200) and is sold by Ted Pella. The last price I have is
about $3000. Another consideration is the size of the chamber. If you run
only a few specimens you might want a smaller one.

Hope this helps, John (and others),

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: kgb-at-gwd.dsto.gov.au (Ken Byfield)
Date: Fri, 5 Nov 1993 14:32:55 +0500
Subject: Image storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9311050501.AA17438-at-spitfire.dsto.gov.au}


If I may add my two bits worth concerning Kodak's Photo CD system:
(- and aplogies in advance if this information has already been covered in
previous discussions)
KODAK offer an Product Information Kit re Photo CD and their image
processing covering
o Pro Photo CD (6000 by 4000 pixels of resolution)
o Lite Box image library system
o Photo CD Catalogue (6000 image storage)
o Portfolio Photo CD (variable resolution from base allowing 800
images or 1 hour of sound, or any combination in between)
and more

Hope that was of some use to someone!
Cheers,




Ken BYFIELD




From: kgb-at-gwd.dsto.gov.au (Ken Byfield)
Date: Fri, 5 Nov 1993 14:32:55 +0500
Subject: Image storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
Phone : 61 8 259 5326
FAX : 61 8 259 6964

Postal Address:-

Bldg. 42 LABS Area
GWD, DSTO Salisbury,
P.O. Box 1500 Salisbury
South Australia 5108

-------------------------------------------
):-) (-:(





From: bnhirsch-at-dapsas1.weizmann.ac.il (David L. Hirschberg)
Date: Fri, 5 Nov 1993 16:26:27 +0200
Subject: Image storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9311051426.AA14036-at-dapsas1.weizmann.ac.il}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

How do you unsubscribe from this list? I am going to be away for 6 weeks
and want to stop the flow of mail.

Listserv-at--at-anlemc.msd.anl.gov

ignors my requests.

-d

=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=
David L. Hirschberg bnhirsch-at-dapsas1.weizmann.ac.il
Department of Neurobiology (972) (0)834-2127 (0)834-2412 work
Weizmann Institute of Science (972) 847-4805 home
Rehovot 76100 Israel (972) 834-4131 fax







From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Fri, 5 Nov 1993 08:48:25 -0700
Subject: Re: Critical point drying

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} On Thu, 4 Nov 1993 04:29:55 +0200, John Mansfield wrote:
}
} } Date:11/3/93
} } NC EMAL
} } Subject: Critical Point Dryer
} } I have a colleague who is in the market for a critical point drying system.
} } Not knowing anything about such things or exactly how they work, I want to ask
} } if there is a manufacturer or model that is preferred by the majority of
} } people
} } and if it is possible to dry material that is wet with a liquid other than
} } water.
} } Any help or info would be appreciated.
} } Thanks.
} } John Mansfield
}
} The idea of critical point drying is to change the liquid to a suitable
} liquified gas, which has a low enough critical point to change directly
} fron liquid phase to gas phase. It does not matter wether the change is
} made directly or via a series of liquids. The only thing which matters is
} to maintain the sample as intact as possible while changing the liquids. So
} all you need to know is the solubility of different liquids in each other
} and finally in the liquified gas.
}
} Hopefully this helps, others can give you more exact advice.
}
} Best regards,
}
} Jouko Maki

The most common transition fluid used these days is liquid carbon dioxide.
Specimens are typically dehydrated through a graded series of organic
solvent, ethanol or acetone, then placed in the cooled critical point dryer
(CPD). Ethanol and acetone are both miscible with CO2. Once the specimens
are in the chamber, liquid CO2 is introduced, and, through several
exchanges (filling and draining with CO2 under pressure from the tank), the
organic solvent is completely replaced by CO2.

The chamber is kept cool to keep the CO2 in liquid phase. After exchange
of liquids, the temperature is increased and the pressure inside the vessel
increases also. Once the critical point is passed, 1073 psi and 31.1 deg C
for CO2, the vessel is slowly vented. After atmospheric pressure is
reached, the chamber is opened and the specimens removed. They are warm
and dry, and they tend to rehydrate pretty fast, so having a dessicator
handy is a good idea.

The typical CPD device is a high pressure bomb that has valves (3) for
introducing liquid CO2 and venting liquid (bottom of chamber) and gas (top
of chamber). There is a mechanism for heating and cooling, too. One
temperature control system has electronic heating and cooling devices
(can't think of the name now) and sensors for temp control. These can be
automated. The other kind, which I prefer, has a water jacket around the
pressure vessel. You run ice water in to cool and warm/hot water to heat.
You have to keep an eye on the chamber to make sure you don't heat too
fast. But I watch the chamber, even when it's automated.

I like the water controled units because (1) they are cheaper (2) there are
fewer parts to go bad (3) they are smaller. The one I use now is made by
Polaron (model 3200) and is sold by Ted Pella. The last price I have is
about $3000. Another consideration is the size of the chamber. If you run
only a few specimens you might want a smaller one.

Hope this helps, John (and others),

John chandler-at-lamar.ColoState.EDU Fort Collins, CO





From: kayton-at-ohsu.edu (Robert Kayton,MAC,CROET)
Date: Tue Nov 9 10:25:29 PST 1993
Subject: E.M. Long term storage of tissues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I need several options for storage of fixed tissues that might be used later
for E.M. Currently I store tissues in the original fixative. Any other
options would be appreciated.

Return E-mail address kayton-at-ohsu.edu




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 10 Nov 1993 16:30:44 -0600 (CST)
Subject: SEM: Environmental

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: llsutter-at-mtu.edu (Larry Sutter)
Date: Wed, 10 Nov 1993 16:55:32 -0500
Subject: Re: Undeliverable Mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I tried posting a message to the list server and got the following reply.
Any thoughts?

} Bad address -- {MICROLISTUSA}
} Error --
} %MAIL-E-NOSUCHUSR, no such user MICROLISTUSA
}
} Start of returned message
}
} Date: Wed, 10 Nov 1993 16:47:40 -0500
} Message-Id: {199311102147.AA09853-at-mtu.edu}
} To: microscopy-at-anlemc.msd.anl.gov
} From: llsutter-at-mtu.edu (Larry Sutter)
} Subject: ESEM
}
} We are in the initial stages of evaluating an environmental SEM.
} I have heard from some people that a field emission SEM with a cryo stage
} will provide very comparable performance. This is an important distinction
} to make early because we definitely have application for a field emission
} SEM for work on photo catalysts, polymers, and materials research. The
} primary interest in an environmental SEM is for examining sludges and
} soils. If this work can be done on a FE/SEM with a cryo stage, our opinion
} is that the FE/SEM would ultimately be more versatile.
}
} I would greatly appreciate any input on this matter especially from
} (but not limited to) people that have used both types of instruments.
}
} Thanks...
}
}
} End of returned message

Thanks

Larry Sutter

Scattering Events R' Us







From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 10 Nov 1993 17:00:20 -0600 (CST)
Subject: NEWS OF RMS MEETINGS FOR 1994

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopy Subscribers:

In case you haven't noticed there have been some glitches
in the system the last few days. The most common problem
reported was either NOMAIL or BAD Addresses. I think
the system is fixed, however, I'm not 100% sure. The mailing
list has grown to ~ 400 users now and I've had to reconfigure
the system to handle 2 groups USA vs. the rest of the World.
Because of this it was necessary to manually edit system subscription
files. I hope no one was erroneously deleted, however, some of you
may be receiving 2 copies of messages. In a 2nd mailing
I will attempt to identify duplicate addresses. For the
time being please excuse the headaches, multiple copies
and various test messages.

Please report any errors directly to me at

zaluzec-at-anlemc.msd.anl.gov

Nestor Zaluzec
ANL EM Center

=======================
BTW here is a message which I believe only reached
part of the mailing list. It was originally sent out on Nov. 7th...

========================






ROYAL MICROSCOPICAL SOCIETY MEETINGS CALENDAR

17 March 1994
ANNUAL IMMUNOCYTOCHEMISTRY MEETING

29 March 1994
ANNUAL LIGHT MICROSCOPY MEETING, London

11 - 13 April 1994
MICROSCOPY OF COMPOSITE MATERIALS II, University of Oxford

13 April 1994
ANNUAL CYTOMETRY AND HISTOCHEMISTRY MEETING,
Birmingham
Organizers: Dr Gillian Lawrence and Dr B. K Shenton

Outline Programme
10.30 am Joint Session, Cytometry and Histochemistry in Pathology
2.00 pm Parallel Sessions
A Histochemistry - Viviane Maggi Prize
Competition
B Cytometry
3.45 pm Pearse Prize Lecture - Professor S. J. Holt
4.30 pm Bingo Meyer Memorial Lecture

Contributed papers are requested for the joint session, the cytometry session
and the Viviane Maggi Prize Competition for beginners. The closing date for
receipt of abstracts will be 31 January 1994.


18 - 22 April 1994
EM SPRING SCHOOL, University of Sheffield

28 June 1994
FLOW CYTOMETRIC IMMUNOPHENOTYPING OF LEUKAEMIAS
AND LYMPHOMAS MEETING, London

27 - 30 June 1994
COMPUTERS IN MICROSCOPY COURSE, University of Cambridge

13 - 14 July 1994
BENCH-TOP FLOW CYTOMETRY WORKSHOP, Newcastle University

17 - 22 July 1994
LM SUMMER SCHOOL, University of Leeds

5 - 9 September 1994
IMMUNOCYTOCHEMISTRY COURSE, Oxford Polytechnic

12 - 15 September 1994
MICRO 94 EXHIBITION AND CONFERENCE, London

12 September 1994
AGM AND PRESIDENTIAL LECTURE, London

19 - 23 September 1994
FLOW CYTOMETRY COURSE, University of Cambridge

October 1994
MICROSCOPY OF CATALYSIS, Joint Royal Microscopical Society/Royal
Society of Chemistry Meeting, London

October 1994
FUNCTION TESTING BY FLOW CYTOMETRY WORKSHOP

November 1994
ULTRASTRUCTURAL IMMUNOCYTOCHEMISTRY COURSE, Sutton

Details of all meetings are available from the Royal Microscopical Society,
37-38 St Clements, Oxford, OX4 1AJ, United Kingdom. Telephone +44
(0)865 248768, fax +44 (0)865 791237, email RMS-at-UK.AC.OX.VAX.



MICRO 94 FIRST CIRCULAR


MICRO 94, International Microscopy and Image Analysis Conference and
Exhibition
12-15 September 1994

Earls Court Park Inn International, Lillie Road, London, SW6


Description of Conference

The Conference will consist of tutorial and review lectures, technical lectures
organized by the trade, and posters. Experts in the field of instrumentation,
analytical methodology in physics and materials sciences, cytometry,
cytochemistry, cytology, image analysis and image processing are invited to
give the lectures.

Each speaker will provide an overview of the particular topic in question and
ample time for discussion will be provided. Seven half-day sessions of tutorial
and review lectures and six half-day sessions of technical lectures will be
organized.

Full length papers of the invited tutorial lectures will be considered for
publication in the Journal of Microscopy.

Contributed papers are welcomed and will appear in poster sessions. Abstracts
will appear in the Proceedings of the Royal Microscopical Society.

The Physiological Society and the Electron Microscopy and Analysis Group
of the Institute of Physics (EMAG) will organize separate lectures within the
scope of the conference.

Outline Scientific Programme

At the MICRO 94 Conference and Exhibition, emphasis will be placed on
Image Processing and Analysis. The launch of a new Special Interest Group
within the Royal Microscopical Society will also be taking place.

Review and Tutorial Lecture topics are expected to include:-

Image processing and analysis
Computer manipulations of images
Scanning probe microscopy in materials sciences and in life sciences
Near field optical microscopy
3-D microscopy and reconstruction
Confocal scanning laser microscopy
Materials
Microscopy of composite materials
Microscopy of electronic materials
Nanostructures
Cytometry and Cytochemistry
Study of cell proliferation with flow cytometry
Tumour markers
Probes for in situ hybridisation
Probes for immunocytochemistry (p53, BrdU, Ki67, PCNA)
Molecular probes for analysis of living cells (co-sponsored by The
Physiological Society)
Proteases in pathological processes
Invasion and metastasis of cancer cells
Arthritis and rheumatism
Infections
Environmental SEM

Technical Lectures will be organized by Exhibitors to act as a bridge between
the specialized review lectures and the equipment being exhibited.

Furthermore, on the Monday, a special session on how to use the light
microscope will take place.


Exhibition

An Exhibition of the latest microscopes, ancillary instrumentation and
technology, of material needed for preparation of microscopic specimens
(fluorescent probes, antibodies and dyes) and soft and hardware for image
processing and image analysis will be held at the Earls Court Park Inn
International. Admission to the exhibition is free by conference badge or
exhibition only badge which will be obtainable at the registration desk.
Exhibition tickets will also be available well before the exhibition takes place.

There will be approximately 50 companies exhibiting, providing an extensive
range of equipment for you to view.

If you would like further details of the MICRO 94 Conference and
Exhibition, please contact the Royal Microscopical Society, 37-38 St
Clements, Oxford, OX4 1AJ, United Kingdom. Telephone +44 (0)865
248768, fax +44 (0)865 791237, email RMS-at-UK.AC.OX.VAX. Conference
enquiries should be addressed to Miss Karen Hale, Exhibition enquiries
should be addressed to Mrs Allison Winton.

-------------------end of message----------------------





From: Jouko K. Maki :      jokamaki-at-utu.fi
Date: Thu, 11 Nov 1993 08:28:14 +0200
Subject: Email address of the RMS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Details of all meetings are available from the Royal Microscopical Society,
} 37-38 St Clements, Oxford, OX4 1AJ, United Kingdom. Telephone +44
} (0)865 248768, fax +44 (0)865 791237, email RMS-at-UK.AC.OX.VAX.


I believe, the Email address of the RMS is written in the wrong order. To
my knowledge, it should read
rms-at-vax.ox.ac.uk

Please, correct me if I am wrong.


Jouko Maki

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Jouko K. Maki Navigare necesse est...
Laboratory Manager, Ph.D.
Laboratory of Electron Microscopy Kiinamyllynkatu 10 FIN-20520 TURKU FINLAND
University of Turku Tel.: + 358 21 633 7318
INTERNET: jouko.maki-at-utu.fi FAX.: + 358 21 633 7380




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Thu, 11 Nov 1993 21:45:36 -0600 (CST)
Subject: Listserver Problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Via: uk.ac.rpms; Thu, 11 Nov 1993 07:09:38 +0000
Via: mpcc3.rpms.ac.uk; Thu, 11 Nov 93 08:13:01 GMT
anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-ml.wpafb.af.mil, waheeschen-at-dow.com,
winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu, DUNLAP-at-UTKVX.UTCC.UTK.EDU,
tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM,
dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
CC: ZALUZEC-at-anlemc.msd.anl.gov
X-Vmsmail-To: -at-micro
X-Vmsmail-Cc: ZALUZEC
Message-Id: {931111222108.2020031a-at-anlemc.msd.anl.gov}

MAIL FROM: {dlc-at-owlnet.rice.edu}
RCPT TO: {MICROSCOPYLISTUSA}
ARRIVAL_TIME: 11-NOV-1993 09:35:01
anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-sparky.ml.wpafb.af.mil,
waheeschen-at-dow.com, winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu, DUNLAP-at-UTKVX.UTCC.UTK.EDU,
tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM,
dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
CC: ZALUZEC-at-anlemc.msd.anl.gov
X-Vmsmail-To: -at-MICRO
X-Vmsmail-Cc: ZALUZEC
Message-Id: {931111214536.2020031a-at-anlemc.msd.anl.gov}

Microscopy Subscribers

The "small" problem I thought I solved earlier has
turned out to be just the tip of an iceberg. The
server will keep running however, it will be
intermittent while I try to debug the system.

Part of the problem appears to be in "Multinet" the Internet
gateway I'm using here at ANL. Some of the
messages you will be receiving in the next
few days will have double headers while I
sort out the works. The first header will likely
always show my name, while the second one
will be the true sender.

Please keep me (Zaluzec-at-ANLEMC.MSD.ANL.GOV )informed of
problems that you see so that I can continue debugging.

In the interim just keep posting messages to

Microscopy-at-ANLEMC.MSD.ANL.GOV

and I'll sort the confusion at this end.

Sorry Nestor :-(
Zaluzec-at-ANLEMC.MSD.ANL.GOV





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 9 Nov 1993 22:37:48 -0600 (CST)
Subject: EM: Long Term Tissue Storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-ml.wpafb.af.mil, waheeschen-at-dow.com,
winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu, DUNLAP-at-UTKVX.UTCC.UTK.EDU,
tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM,
dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
CC: ZALUZEC-at-anlemc.msd.anl.gov
X-Vmsmail-To: -at-micro
X-Vmsmail-Cc: ZALUZEC
Message-Id: {931111215718.2020031a-at-anlemc.msd.anl.gov}

MAIL FROM: {simon-at-sbic.cbio.pitt.edu}
RCPT TO: {MICROSCOPYLISTUSA}
ARRIVAL_TIME: 11-NOV-1993 19:26:04
anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-ml.wpafb.af.mil, waheeschen-at-dow.com,
winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu, DUNLAP-at-UTKVX.UTCC.UTK.EDU,
tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM,
dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
CC: ZALUZEC-at-anlemc.msd.anl.gov
X-Vmsmail-To: -at-micro
X-Vmsmail-Cc: ZALUZEC
Message-Id: {931111223242.2020031a-at-anlemc.msd.anl.gov}

MAIL FROM: {ZALUZEC}
RCPT TO: {MICROSCOPYLIST}
ARRIVAL_TIME: 9-NOV-1993 22:37:57


} Kayton asks:
I need several options for storage of fixed tissues that might be used later
for E.M. Currently I store tissues in the original fixative. Any other
options would be appreciated.

======
What is the time scale you need to store your samples?
Are they already EM (?TEM/SEM?) specimens?

Nestor Z.
ANL EM Center




From: richard crang :      rcrang-at-vmd.cso.uiuc.edu
Date: Thu, 11 Nov 1993 14:44:12 -0600 (CST)
Subject: LM - Need help on staining epoxy/Spurr embedded plant tissues_

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-ml.wpafb.af.mil, waheeschen-at-dow.com,
winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu, DUNLAP-at-UTKVX.UTCC.UTK.EDU,
tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM,
dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
CC: ZALUZEC-at-anlemc.msd.anl.gov
X-Vmsmail-To: -at-micro
X-Vmsmail-Cc: ZALUZEC
Message-Id: {931111221716.2020031a-at-anlemc.msd.anl.gov}

MAIL FROM: {ZALUZEC}
RCPT TO: {MICROSCOPYLISTUSA}
ARRIVAL_TIME: 11-NOV-1993 14:44:27


48 hr. post-germination pea root tips and segments through differentiation
zone were fixed in 2.5% glutaraldehyde in 0.15 M cacodylate buffer (pH 6.8),
followed by 2% OsO4 in same buffer, then dehydrated and embedded in a resin
mixture of equal parts epon (substitute=Epox 812): Spurr. 1-2 micron
sections do not stain with 1% toulidine blue with varied times and temps.
Am interested in good general stain which will also reveal mitotic
figures__ANY SUGGESTIONS?




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 12 Nov 1993 9:36:49 -0600 (CST)
Subject: LM - Need help on staining epoxy/Spurr embedded plant tissues_

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-sparky.ml.wpafb.af.mil,
waheeschen-at-dow.com, winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-sparky.ml.wpafb.af.mil,
jfc3348-at-u.cc.utah.edu, stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu,
DUNLAP-at-UTKVX.UTCC.UTK.EDU, tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu,
DTHAUS-at-VNET.IBM.COM, dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV,
jkelly-at-enh.nist.gov, morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu,
ssun-at-helix.nih.gov, cooper-at-ipl.rpi.edu, microscopy-at-di.com,
preynold-at-itsmail1.hamilton.edu
X-Vmsmail-To: -at-MICRO
Message-Id: {931112093649.20200798-at-anlemc.msd.anl.gov}




From: NAME \ Greg Erdos, ICBR EM Core Lab.\ :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Wed, 10 Nov 1993 09:18:41 -0500 (EST)
Subject: Tissue Storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-ml.wpafb.af.mil, waheeschen-at-dow.com,
winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu, DUNLAP-at-UTKVX.UTCC.UTK.EDU,
tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM,
dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
CC: ZALUZEC-at-anlemc.msd.anl.gov
X-Vmsmail-To: -at-micro
X-Vmsmail-Cc: ZALUZEC
Message-Id: {931111221124.2020031a-at-anlemc.msd.anl.gov}

MAIL FROM: {ZALUZEC}
RCPT TO: {MICROSCOPYLISTUSA}
ARRIVAL_TIME: 11-NOV-1993 14:48:22


MAIL FROM: {dlc-at-owlnet.rice.edu}
RCPT TO: {MICROSCOPYLISTUSA}
ARRIVAL_TIME: 11-NOV-1993 09:47:07
anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu, baskindg-at-u.washington.edu,
Baginski-at-USUHSB.USUHS.MIL, BEGGCA%ML%WPAFB-at-MLGATE.ML.WPAFB.AF.MIL,
Bradley-at-ANLEMC.MSD.ANL.GOV, bbug1-at-swarthmore.edu,
bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org, Borisy-at-macc.wisc.edu,
cammer-at-aecom.yu.edu, carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov, clancy-at-mat.mte.ncsu.edu,
colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov, csencsits-at-anlemc.msd.anl.gov,
disko-at-hal.erenj.com, dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu,
dolber-at-cs.duke.edu, e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu, giliciag-at-ttown.apci.com,
gilkey-at-biosci.arizona.edu, Gilroy-at-shiva.psu.edu, glenmac-at-u.washington.edu,
gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov, gwerdos-at-gnv.ifas.ufl.edu,
Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu, HECUB-at-ttacs1.ttu.edu,
herro001-at-maroon.tc.umn.edu, hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu, lawrence.bottomley-at-chemistry.gatech.edu,
leapman-at-helix.nih.gov, llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu,
jcrum-at-sunstroke.sdsu.edu, JEANNE_BARKER-at-MERCK.COM,
JMARDINLY-at-MATTEC.INTEL.COM, john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu, morilak-at-cmgm.stanford.edu,
mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu, naresh-at-funky.mm.uky.edu,
noah-at-helix.nih.gov, pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu, tah-at-med.pitt.edu,
tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov, TJOSEPH-at-VNET.IBM.COM,
tonygr-at-mit.edu, trimblea-at-zoology.washington.edu, tsu-at-cae.wisc.edu,
twesten-at-uimrl7.mrl.uiuc.edu, Walcksd-at-ml.wpafb.af.mil, waheeschen-at-dow.com,
winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU, yang-at-msc.cornell.edu,
Zeissler-at-gapnet.nist.gov, MicroArchive-at-anlemc.msd.anl.gov,
yu_lian-at-lilly.com, dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu, ag14-at-andrew.cmu.edu,
dave-at-mando.mit.edu, davilla-at-marimba.cellbio.duke.edu,
dale_batchelor-at-ncsu.edu, KEEBLE-at-mtu.edu, tfoecke-at-nist.gov,
knecht-at-uconnvm.uconn.edu, jester-at-crnjjsgi.swmed.edu,
spector-at-surface.ucsb.edu, bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov, chia-at-msc.cornell.edu,
jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM, sfh-at-rml.niaid.nih.gov,
SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org, cfg-at-rml.niaid.nih.gov,
mossant-at-ducvax.auburn.edu, vandezande-at-inland.com,
tatistcheff-at-milkwy.enet.dec.com, cllin-at-i1.chen.umn.edu,
Pguthrie-at-vines.ColoState.EDU, marko-at-hvem.ph.albany.edu,
adam-at-riscy.ucsb.edu, QJXNJ21-at-MEMRQA.SPS.MOT.COM,
microscopy-at-cheme.washington.edu, GOh.313-at-postman.gsfc.nasa.gov,
jmichael-at-vnet.ibm.com, bbanke-at-btvmanvm.vnet.ibm.com,
mwfolsom-at-hydra.unm.edu, wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu,
mmhenson-at-med.unc.edu, walcksd-at-ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu, DUNLAP-at-UTKVX.UTCC.UTK.EDU,
tm11-at-umail.umd.edu, sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM,
dfalcon-at-vnet.ibm.com, Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
CC: ZALUZEC-at-anlemc.msd.anl.gov
X-Vmsmail-To: -at-micro
X-Vmsmail-Cc: ZALUZEC
Message-Id: {931111222842.2020031a-at-anlemc.msd.anl.gov}

MAIL FROM: {GWERDOS-at-gnv.ifas.ufl.edu}
RCPT TO: {MICROSCOPYLIST}
ARRIVAL_TIME: 11-NOV-1993 08:20:35


Robert Kayton asked about fixed tissue storage. We generally advise the use
of Trumps fixative as it has bee repoted good for this purpose. (McDowell &
Trump, 1976, Arch. Pathol Lab. Med. 100:405)

Last year there was a paper by Sopsi et al. (Ultrastructural Pathol. 16:351)
that reported good results storing fixed tissue frozen at -70 C after cryo-
protection with sucrose (10% I think). They report good immunoreactivity after
2 years and it looked pretty good.

Greg Erdos
Univ. of Florida
gwerdos-at-gnv.ifas.ufl.edu




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 12 Nov 1993 16:22:53 -0600 (CST)
Subject: Test of Fixed Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



This is a test of the reconfigured software.
please ignore this message. (I hope things are fixed now!)

Nestor Z.




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 12 Nov 1993 19:18:15 -0600 (CST)
Subject: ... and the problem was

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Hooray!!!!!!!!!!!

It's solved. The problem now definitely appears cured.

You can delete the rest of this message now, unless you
want to know what happened.

========================

Symptom: Mailserver lockup, the batch server takes } 2 days
to deliver mail that previously took 1/2 hour to process.

Cause:

Ultimately the mailserver lockup was caused by a faulty
DNS (Domain Name Server) Address. The subscriber provided
an address which checked out initially, remember that
the instructions ask you to verify your address! However, after
being added to the subscription list, something went
haywire with one host node. In effect the mailserver would
spend ~ 1-2 days searching & eventually finding the location of the
subscriber. Unfortunately the way this software runs
each subscriber is accessed sequentially and thus the
mail would "appear" to hang doing nothing for a day
(or more) while this one host is identified using god
know how many secondary, teriary, DNS's in effect
bouncing it's way around internet.


Solution:
I've removed the problem address from the
list and will contact the user seperately. Since the
mail actually makes it through to the "lost" host
I had no direct traceback feature to locate him and
it took about a week to find. Some of you may not
have noticed a problem since your subscription would
have been entered and cleared before the problem one.
In effect the only way to find the problem was to
manually test the addresses, something I do not
recommend that anyone does.

So... sorry that some of you had to put up with:
1.) multiple copies of repeated messages:
2.) excessively long headers with 1/2 the worlds
email addresses
3.) multiple test messages

Many of you will not have received all of the
messages posted to the listserver during this last
week. If I see one that looks very important that
in my infinite wisdom should be reposted I'll do so,
otherwise it will just sit until the topic
becomes posted again by the original author.
I have an archive of everything and eventually
will setup some way for anyone to search that archive
but I've got alot of catching up to do and this is
extremely low on my priority list of things to do.


That's all Folks! :-)

Nestor Z.
ANL EM Center

Some how I just knew this listserver would become masochistic






From: Sally Stowe :      STOWE-at-rsbs-central.anu.edu.au
Date: Sat, 13 Nov 1993 08:52:43 EST10
Subject: TEM-LaB6 in H600

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




In our standard H600 TEM, we have been getting comparable tungsten
filament life to that in some of our other machines configured with
sputter pumps, for use with LaB6 cathodes. Emboldened by this, we
are fitting a Lab6 to this system to see what happens - anybody out
there who has already done this, or has any comments? The cathode we
have is a Denka, although we are generally moving to Kimball.

----------------------------------------------------------------------
Sally Stowe |
Facility Coordinator
ANU Electron Microscopy Unit |
Australian National University
Canberra, AUSTRALIA |
Ph 61 6 249 2743
FAX 61 6 249 4891 |
Email stowe-at-rsbs-central.anu.edu.au
-------------------------------------|--------------------------------
-





From: John.F.Mansfield-at-umich.edu (John F. Mansfield)
Date: Thu, 11 Nov 1993 11:22:26 -0500
Subject: Any interest in a microscopy newsgroup?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu

I just sent off a posting suggesting the creation of a micrscopy newsgroup on
the UseNet News system.

It's in the following newsgroups:
sci.chem,sci.engr.chem,sci.geo.geology,sci.polymers,sci.materials,sci.misc,sci.p
hysics,sci.research,sci.techniques.xtallography,sci.optics,sci.archaeology,sci.
engr.mech,sci.geo.geology,sci.med

The following text is the full body of the posting, I encourage comments from
the microscopy and imaging community. Please feel free to add to the necessary
areas of interest, I will post follwo-ups when I have some response. I only
posted this about two hours ago and I have had two messages of support already.

Path: caen!jfm.sprl.umich.edu!user
hysics,sci.research,sci.techniques.xtallography,sci.optics,sci.archaeology,sci.
engr.mech,sci.geo.geology,sci.med

This post could equally well be titled:
pre-RFD : sci.techniques.microscopy

Following is a proposed RFD (request for discussion) to create a
microscopy newsgroup on the internet

Would potential users of this proposed newsgroup like to give their
opinions on this (YES or NO) either by following on to this post (which
should appear in sci.materials) or as email to
(John.F.Mansfield-at-umich.edu).

Would people also consider passing this proposal onto colleagues who may
be interested in participating in a microscopy newsgroup. Potential
users of this newsgroup might like to consider adding their names to the
RFD as co-proposers. It is hoped that at least one-hundred people can
give support to this newsgroup before the RFD is officially posted to
news.groups so as to show there is decent support for a microscopy
newsgroup on the internet. If it looks like this will take a while to
do, summaries of responses (who will remain anonymous) will be posted
regularly to keep people informed what has been happening.

The reason for the name sci.techniques.microscopy is that the Usenet
administrators have created the "techniques" hierarchy for newsgroups
such as microscopy. Going with this name has the best hope of
avoiding hair splitting arguments on newsgroup "names" that seem to
plague the creation of other science newsgroups. The above name does
not affect the aims or the charter of what this microscopy newsgroup
is trying to achieve.

Please note if a particular topic of microscopy becomes very popular
or there is a perceived need to create a subset of this newsgroup such
as "SEM", "AFM" or "TEM", etc - it is relatively easy to create a

sci.techniques.microscopy.sem
or sci.techniques.microscopy.tem, etc,

providing the support for it is there.

=========================================================

Pre - Request For Discussion (RFD) of sci.techniques.microscopy
(unmoderated).

Proposers :- John Mansfield (John_Mansfield-at-mse.engin.umich.edu)
Lachlan Cranswick (lachlan-at-dmp.csiro.au)

Please note that this proposed newsgroup is intended to
be an open forum for discussion of microscopy. Thus
relevant topics for this newsgroup should only be limited
to what the participants in this proposed newsgroup
regard as microscopy.

--------------------------------------------

We propose a new unmoderated newsgroup called
sci.techniques.microscopy. The main aim of
sci.techniques.microscopy is to provide an open forum for the
discussion of microscopy and related fields on the internet.

(Could people reading this post mention it to colleagues who
might be interested and benefit from a microscopy group
but not normally use the internet newsgroups.)

Proposed Name of Group
-------------------------
sci.techniques.microscopy (unmoderated)

Purpose of New Newsgroup
---------------------------

The purpose of sci.techniques.microscopy is to provide an
open discussion forum for the microscopy community on
the internet. The newsgroups allow the rapid and timely discussion of
opinions and information that would take months or years (or not at all)
on conventional paper journals.

Topics for discussion would include :-
---------------------------------------
Optical Microscopy
Confocal Microscopy
Scanning Probe/Force Microscopy (formerly Atomic Force Microscopy) - SPM,
SFM, AFM
Scanning Tunnelling Microscopy - STM
Scanning Electron Microscopy - SEM
Transmission Electron Microscopy - TEM
High Resolution Electron Microscopy - HREM
Analytical Electron Microscopy - AEM
Scanning Transmission Electron Microscopy - STEM
High Voltage Electron Microscopy - HVEM
X-ray Energy Dispersive Spectroscopy - XEDS
Electron Energy Loss Spectroscopy - EELS
Diffraction Contrast Imaging
Phase Contrast Imaging
Selected Area Electron Diffraction - SAED or SAD
Convergent Beam Electron Diffraction - CBED
Image Filtering
Specimen Preparation (Electropolishing, Ion Milling, Ultramicrotoming,
etc.)
Software
Data formats
Databases
Hardware/Equipment - specs, opinions, etc.
Applications
Announcements/reviews of papers/conferences.
Preparation techniques.
Non-ambient techniques
General Discussion/opinions/questions.
Positions vacant

Anything else that is relevant to microscopy in general.

---------------------------------------------------------------

What is the Process of Creating a Newsgroup?
--------------------------------------------

} (a) RFD: Discussion, i.e., public hearing to take place
in the newsgroup news.groups for approximately one month
(b) CFV: Call for votes (the voting period will be about 25 days)
(c) Counting of votes and public display of votes
(d) Announcement of new newsgroup

(a)--} (b) assumes no major disagreements about this newsgroup during
discussion. (c)--} (d) assumes that the vote is favourable., i.e.,
Y } N+100 .and. Y } (2/3)(Y+N)
Y being the number of YES votes, N being the number of NO votes for
the creation of the proposed newsgroup.
=============================================

Please comment/criticise/suggest by followons to this post or email to
John.F.Mansfield-at-umich.edu


John Mansfield, North Campus Electron Microbeam Analysis Laboratory,
2455 Hayward, Ann Arbor, Michigan 48109-2143. Ph: (313)-936-3352.
__________
YYURYYUBICURYY4ME.







From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sat, 13 Nov 1993 22:39:25 -0600 (CST)
Subject: TEM: FREE SPARE PARTS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sat, 13 Nov 1993 22:39:25 -0600 (CST)
Subject: TEM: FREE SPARE PARTS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hitachi HU-11E Parts for free:
If you are nursing along an HU-11E TEM, and would
like some spares for it, please contact me directly
(please don't respond to the list!)
--Forepumps and DP are spoken for
--Most other parts go to anyone who can use them. I'll
ship light stuff inside the continental US for free; you'll
have to pay shipping on the heavy stuff.

If I haven't heard from someone within two weeks, it
goes in the dumpster. (TBAITW; as in "too big and in
the way").

Julian Smith III
Winthrop University
803-323-2111 (Vox)
803-323-2246 (Fax)




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sat, 13 Nov 1993 23:01:51 -0600 (CST)
Subject: TEM:LaB6 in Hitachi 600

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




} Sally Stowe asks about swapping W/LaB6 in a H600.

Sorry but I'm geting old and forget full so
I can't remember the details of an H600, but, the appropriate
question to ask is why are you attempting to put LaB6 in
the H600? not if it will work! The LaB6 filaments when
operated properly and under *CORRECT* conditions will
give you a greater Brightness and hence more electrons
in electron-optically equivalent probes. Thus for small
probe/analytical work or for imaging conditions which
will improve with a more coherent source (HREM) and
in good vacuum systems they work very well.
It will improve your uscope work assuming that you are short
on electrons in the configuration your are setting up
your electron probe (the argument applies to both AEM
and/or HREM modes), however, if you poison the filament by using
it in a poor vacuum system (i.e form an oxide or carbide
on the surface), then the work function will decrease and
you will actually get worse performance than a standard
W filament and at ten to twenty times the cost of W. So
why do you want to do this?

Assuming you have appropriate reasons and a good vacuum
(and students who do not dump the vacuum system to
air with a hot filament running), then the next question to
ask is your uscope equipped with a filament power supply that
can be modified to drive the tip?

Just my 2 cents worth.


Nestor Z.
ANL EM Center




From: Sally Stowe :      STOWE-at-rsbs-central.anu.edu.au
Date: Mon, 15 Nov 1993 10:00:59 EST10
Subject: TEM:LaB6 in Hitachi H600

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Sally STowe asks about using LaB6 under "tungsten" vacuum
configuration - ie without sputter pump.

} Nestor Zaluzec asks 1. Why?
2. Can the filament power supply be modified to
drive Lab6 tip?

In answer to (1) - we want to get the best resolution we can over
the next few months, and at the moment, on
smallest condenser aperture, we cant make best use of the "spot size"
adjustment because of lack of electrons, although the resolution does
seem to be inproving sharply as far as can be seen. This microscope
does seem to be particularly prone to charge effects or whatever if
the beam is not well spread. Vacuum in the specimen area is 8 x
10-6mbar. The last tungsten filament lasted 330ish hours, so reckon
we have some reasonable users at the moment.Touch wood.
(2) - yes, filament supply can be switched.

----------------------------------------------------------------------
Sally Stowe | Australian National Univ.
Facility Coordinator Canberra, AUSTRALIA
ANU Electron Microscopy Unit | Ph 61 6 249 2743
Email stowe-at-rsbs-central.anu.edu.au | FAX 61 6 249 4891
-------------------------------------|--------------------------------
-





From: VEI011-at-GEEL.DWT.CSIRO.AU (Colin Veitch CSIRO Division of Wool Technology)
Date: Mon, 15 Nov 1993 16:35:04 +1100 (EST)
Subject: AFM - imaging software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We use a Park Scientific Universal BD2 Scanning Probe Microscope,
controlled by a HP 382 workstation. Does anyone know of any software
which will easily (and reliably) extract the images into a format which
can be read by PC's? It is possible to export a TIFF file but the format
appears (according to the operator) to be slighly different to other TIFF
formats.

Any help would be appreciated.

Colin V.

#####################################################################
**********************
* Between the idea *
0------* And the reality *
} ---|--- { * Between the motion *
| * And the act *
/ \ * Falls the Shadow *
_/ \_ * T.S. Eliot *
**********************
Colin Veitch Tel + 61 (0)52 47 2611
CSIRO Division of Wool Technology Tel + 61 (0)52 47 2891 (dir.)
P.O. Box 21 Fax + 61 (0)52 47 2657
BELMONT Vic 3216
Australia

#####################################################################




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 15 Nov 1993 16:31:36 -0600 (CST)
Subject: AFM: Imaging Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Colin V. writes....
} We use a Park Scientific Universal BD2 Scanning Probe Microscope,
} controlled by a HP 382 workstation. Does anyone know of any software
} which will easily (and reliably) extract the images into a format which
} can be read by PC's? It is possible to export a TIFF file but the format


Colin it sounds like Park is at fault. We routine transfer TIFF files
from one computer Vax, Mac, PC, Sun... to another an access TIFF files
with no difficulty. My guess is that Park took short cuts and did not
properly implement TIFF (this happens alot from what I can tell).
You should demand that Park rewrite their code to properly
implement TIFF which is no at Version 6.0. T

There are lots of programs which can read TIFF files and display them
on the PC.

Nestor Z.
ANL EM Center
-------------






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 15 Nov 1993 15:07:38 -0800 (PST)
Subject: Re: LM - Need help on staining epoxy/Spurr embedded plant tissues_

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Full-Name:
Reply-To: minter-at-Kodak.COM
anandamu-at-usc.edu, anis-at-moe.eece.mu.edu, amohan-at-UTKVX.UTCC.UTK.EDU,
APD-at-ATVAX.LORD.COM, as_lea-at-pnl.gov, barnard-at-wadsworth.ph.albany.edu,
bart-at-netcom.com, baskin-at-biosci.mbp.missouri.edu,
baskindg-at-u.washington.edu, Baginski-at-USUHSB.USUHS.MIL,
BEGGCA%ML%WPAFB-at-sparky.ml.wpafb.af.mil, Bradley-at-ANLEMC.MSD.ANL.GOV,
bbug1-at-swarthmore.edu, bergrh-at-memstvx1.memst.edu, bobm-at-coyote.rain.org,
Borisy-at-macc.wisc.edu, cammer-at-aecom.yu.edu,
carl_henderson-at-um.cc.umich.edu, cel1-at-lehigh.edu,
chandler-at-lamar.ColoState.EDU, chapman-at-bigq.enet.dec.com,
chernock-at-vnet.ibm.com, cj_bruckner-lea-at-pnl.gov,
clancy-at-mat.mte.ncsu.edu, colijn.1-at-osu.edu, cook-at-anlemc.msd.anl.gov,
csencsits-at-anlemc.msd.anl.gov, disko-at-hal.erenj.com,
dlmedli-at-ca.sandia.gov, dluchtel-at-u.washington.edu, dolber-at-cs.duke.edu,
e-reuter-at-uiuc.edu, eades-at-uimrl7.mrl.uiuc.edu,
eeffah-at-eniac.seas.upenn.edu, EV2-at-ORNL.GOV, Frethem-at-lenti.med.umn.edu,
George.C.Ruben-at-Dartmouth.EDU, GHB1-at-psuvm.psu.edu,
giliciag-at-ttown.apci.com, gilkey-at-biosci.arizona.edu,
Gilroy-at-shiva.psu.edu, gmei-at-hoh.mbl.edu, gnord-at-mactem.er.usgs.gov,
gwerdos-at-gnv.ifas.ufl.edu, Hall-at-me.udel.edu, Hank-at-Wana.pbrc.Hawaii.edu,
HECUB-at-ttacs1.ttu.edu, herro001-at-maroon.tc.umn.edu,
hillyard-at-msc.cornell.edu, howelld-at-egr.msu.edu,
Hjoan-at-clemson.clemson.edu, inovis-at-rock.concert.net, INGRAM-at-RCC.RTI.ORG,
Kbart-at-hamilton.edu, kcherrey-at-physics.berkeley.edu,
king-at-bioscience.biology.utah.edu,
lawrence.bottomley-at-chemistry.gatech.edu, leapman-at-helix.nih.gov,
llsutter-at-mtu.edu, jan-at-chaos.uchicago.edu, jcrum-at-sunstroke.sdsu.edu,
JEANNE_BARKER-at-MERCK.COM, JMARDINLY-at-MATTEC.INTEL.COM,
john.f.mansfield-at-umich.edu, john_russ-at-ncsu.edu,
jpbrody-at-phoenix.princeton.edu, kayton-at-ohsu.edu,
kordesch-at-helios.phy.ohiou.edu, mallam-at-maroon.tc.umn.edu,
MAUGEL-at-zool.umd.edu, mcpetri-at-anl.gov, mdhouse-at-mendel.berkeley.edu,
MECAVALERI-at-MMM.COM, me-at-leidecker.gsfc.nasa.gov, mgarment-at-macc.wisc.edu,
microscopy-at-noran.com, MILLER.MAHLON-at-igate.abbott.com, minter-at-kodak.com,
mmilgrom-at-cheme.cornell.edu, mohney-at-cae.wisc.edu,
morilak-at-cmgm.stanford.edu, mow-at-vm.cfsan.fda.gov, msachs-at-admin.ogi.edu,
naresh-at-funky.mm.uky.edu, noah-at-helix.nih.gov,
pathology-at-a1.mscf.upenn.edu, plee-at-neep.engr.wisc.edu,
po-at-parmly1.parmly.luc.edu, powell-at-zodiac.rutgers.edu,
provost-at-corl.nbc.upenn.edu, rab-at-fsh.mtu.edu,
RDerro.313-at-postman.gsfc.nasa.gov, redder-at-vnet.ibm.com, rkn-at-ornl.gov,
robinsonp-at-polar.enet.dec.com, ron-anderson-at-vnet.ibm.com,
RMFISHER-at-CARSON.U.WASHINGTON.EDU, rxm-at-atvax.lord.com, SDDEMAGG-at-UCI.EDU,
slking-at-maroon.tc.umn.edu, slimbach-at-macc.wisc.edu, spatel-at-goodyear.com,
spl-at-szechuan.ucsd.edu, smiller-at-umrvmb.umr.edu, SMITHJ-at-Winthrop.edu,
sswami-at-fraser1.eng.ohio-state.edu, STGPMPM-at-geological.unocal.com,
sweeney-at-dionheinz.uchicago.edu, talwar-at-wuche1.wustl.edu,
tah-at-med.pitt.edu, tcizzy-at-casbah.acns.nwu.edu, tgs-at-helix.nih.gov,
TJOSEPH-at-VNET.IBM.COM, tonygr-at-mit.edu, trimblea-at-zoology.washington.edu,
tsu-at-cae.wisc.edu, twesten-at-uimrl7.mrl.uiuc.edu,
Walcksd-at-sparky.ml.wpafb.af.mil, waheeschen-at-dow.com,
winkler-at-gapnet.nist.gov, WOODBA-at-ESVAX.DNET.DUPONT.COM,
wsryu-at-phoenix.princeton.edu, VICKIE-at-MACC.WISC.EDU,
yang-at-msc.cornell.edu, Zeissler-at-gapnet.nist.gov,
MicroArchive-at-anlemc.msd.anl.gov, yu_lian-at-lilly.com,
dbj-at-JAMES.PSYCH.YALE.EDU, ap-at-maelstrom.ucsf.edu,
yprakash-at-siecklab.mayo.edu, huff-at-mcclb0.med.nyu.edu,
ag14-at-andrew.cmu.edu, dave-at-mando.mit.edu,
davilla-at-marimba.cellbio.duke.edu, dale_batchelor-at-ncsu.edu,
KEEBLE-at-mtu.edu, tfoecke-at-nist.gov, knecht-at-uconnvm.uconn.edu,
jester-at-crnjjsgi.swmed.edu, spector-at-surface.ucsb.edu,
bunasawa-at-aludra.usc.edu, tim-at-mae1.engr.ucf.edu,
richard.g.rateick.1-at-nd.edu, dwd-at-rml.niaid.nih.gov,
chia-at-msc.cornell.edu, jbp-at-rtifs2.rti.org, STETSJR-at-TTOWN.APCI.COM,
sfh-at-rml.niaid.nih.gov, SLS13-at-PSUVM.PSU.EDU, cheny-at-mcnc.org,
cfg-at-rml.niaid.nih.gov, mossant-at-ducvax.auburn.edu,
vandezande-at-inland.com, tatistcheff-at-milkwy.enet.dec.com,
cllin-at-i1.chen.umn.edu, Pguthrie-at-vines.ColoState.EDU,
marko-at-hvem.ph.albany.edu, adam-at-riscy.ucsb.edu,
QJXNJ21-at-MEMRQA.SPS.MOT.COM, microscopy-at-cheme.washington.edu,
GOh.313-at-postman.gsfc.nasa.gov, jmichael-at-vnet.ibm.com,
bbanke-at-btvmanvm.vnet.ibm.com, mwfolsom-at-hydra.unm.edu,
wlin-at-aries.scs.uiuc.edu, jstanly-at-mse.ogi.edu, mmhenson-at-med.unc.edu,
walcksd-at-sparky.ml.wpafb.af.mil, jfc3348-at-u.cc.utah.edu,
stuartm-at-maroon.tc.umn.edu, lee-at-engr.wisc.edu,
DUNLAP-at-UTKVX.UTCC.UTK.EDU, tm11-at-umail.umd.edu,
sbarlow-at-sunstroke.sdsu.edu, DTHAUS-at-VNET.IBM.COM, dfalcon-at-vnet.ibm.com,
Zhang-at-BNLX1L.NSLS.BNL.GOV, jkelly-at-enh.nist.gov,
morgan-at-notorious.lbl.gov, fan-at-med.pitt.edu, ssun-at-helix.nih.gov,
cooper-at-ipl.rpi.edu, microscopy-at-di.com, preynold-at-itsmail1.hamilton.edu
In-Reply-To: {931112093649.20200798-at-anlemc.msd.anl.gov}
Message-Id: {Pine.3.87a.9311151538.A21670-0100000-at-carson.u.washington.edu}
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; charset=US-ASCII

Yes, 1.% Toluidine Blue in 1% sodium borate will stain sodium ethoxide
etched Spurr epoxy.

On Fri, 12 Nov 1993, Nestor J. Zaluzec (708)-252-5075, -4964 wrote:

} From: SMTP%"-at-cc1.kuleuven.ac.be:desclinj-at-ulb.ac.be" 12-NOV-1993 02:02:51.52
} To: ZALUZEC
} CC:
} Subj: Re: LM - Need help on staining epoxy/Spurr embedded plant tissues_
}
} From: desclinj-at-ulb.ac.be (Desclin Jean)
} Message-Id: {9311120706.AA00745-at-is1e.vub.ac.be}
} Subject: Re: LM - Need help on staining epoxy/Spurr embedded plant tissues_
} To: ZALUZEC-at-anlemc.msd.anl.gov (Nestor J. Zaluzec)
} Date: Fri, 12 Nov 93 8:06:14 MET
} In-Reply-To: {931111144412.202000ba-at-anlemc.msd.anl.gov} ; from "Nestor J. Zaluzec" at Nov 11, 93 2:44 pm
} X-Mailer: ELM [version 2.3 PL11]
}
} Hello!
} what is the pH of your toluidine blue solution? It might help
} trying out a higher pH than usual (I am afraid I am not familiar
} with plant tissues :-(...).
} If you are not re-using the semi-thin sections afterwards (i.e.
} resectioning for TEM), you might try treating them with NaOH
} dissolved in absolute ethanol (to etch the plastic somewhat), such
} as used prior to some immunohistochemical procedures. I believe
} toluidine blue might work after that.
} Good luck!
} John
}
}
} ***********************************************************
} * Jean C. Desclin (John), Associate Prof. of Histology *
} * Laboratory of Histology - Faculty of Medicine *
} * Brussels Free University (U.L.B.) *
} * e-mail: desclinj-at-ulb.ac.be (internet) *
} * snail mail: route de Lennik 808 *
} * B - 1070 Brussels Belgium *
} ***********************************************************
}





From: VEI011-at-GEEL.DWT.CSIRO.AU (Colin Veitch CSIRO Division of Wool Technology)
Date: Tue, 16 Nov 1993 16:36:35 +1100 (EST)
Subject: AFM - Image software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Many thanks to those who offered information regrding my query. All the
information has been most illuminating (and useful).

Colin V.

#####################################################################
*******************************
* Logic is invincible because *
0------* in order to combat logic it *
} ---|--- { * is necessary to use logic. *
| * P.Boutroux *
/ \ *******************************
_/ \_

Colin Veitch Tel + 61 (0)52 47 2611
CSIRO Division of Wool Technology Tel + 61 (0)52 47 2891 (dir.)
P.O. Box 21 Fax + 61 (0)52 47 2657
BELMONT Vic 3216
Australia

#####################################################################






From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 15 Nov 1993 16:31:36 -0600 (CST)
Subject: AFM: Imaging Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Colin V. writes....
} We use a Park Scientific Universal BD2 Scanning Probe Microscope,
} controlled by a HP 382 workstation. Does anyone know of any software
} which will easily (and reliably) extract the images into a format which
} can be read by PC's? It is possible to export a TIFF file but the format


Colin it sounds like Park is at fault. We routine transfer TIFF files
from one computer Vax, Mac, PC, Sun... to another an access TIFF files
with no difficulty. My guess is that Park took short cuts and did not
properly implement TIFF (this happens alot from what I can tell).
You should demand that Park rewrite their code to properly
implement TIFF which is no at Version 6.0. T

There are lots of programs which can read TIFF files and display them
on the PC.

Nestor Z.
ANL EM Center
-------------






From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Tue, 16 Nov 1993 11:07:51 -0500 (EST)
Subject: infiltration of yeast with spurrs and lr white

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm starting to do gold labeling of protein in yeast cells and I'm having
problems with infiltration. This is my first time in 25 years I've had to
work with yeast. Any helpful hints?

Phil Rutledge
UMBC Biology Dept.
Catonsville, MD 21228
Phone: (410) 455-3582
FAX: (410) 455-3875
"E" Mail: Prutle1-at-umbc.edu

Thanks





From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Tue, 16 Nov 1993 09:45:41 -0800 (PST)
Subject: re: staining spurr

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Yes, 1.% Toluidine Blue in 1% sodium borate will stain sodium ethoxide
etched Spurr epoxy. I've only had toluidine fail in Spurr's if the pH was
off (someone didn't make up the correct borate } . For autoradiography
and nuclear anntigen immunocytochemistry, it can be diluted to 0.1% or
even 0.01% in 1% sodium borate to keep from obscuring the label.

} } If you are not re-using the semi-thin sections afterwards (i.e.
} } resectioning for TEM), you might try treating them with NaOH
} } dissolved in absolute ethanol (to etch the plastic somewhat), such
} } as used prior to some immunohistochemical procedures. I believe
} } toluidine blue might work after that.
} } Good luck!
} } John
} }
} }
} } ***********************************************************
} } * Jean C. Desclin (John), Associate Prof. of Histology *
} } * Laboratory of Histology - Faculty of Medicine *
} } * Brussels Free University (U.L.B.) *
} } * e-mail: desclinj-at-ulb.ac.be (internet) *
} } * snail mail: route de Lennik 808 *
} } * B - 1070 Brussels Belgium *
} } ***********************************************************
} }
}
}





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 15 Nov 1993 16:31:36 -0600 (CST)
Subject: AFM: Imaging Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"Nestor J. Zaluzec (708)-252-50" {ZALUZEC-at-anlemc.msd.anl.gov}
CC: ZALUZEC-at-anlemc.msd.anl.gov

Reply_ RE} AFM: Imaging Software
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
We have a Nanoscope 3 and Digital Instruments have a TIFF export that NIH-Image
doesnt like. When we import the images into image we lose the LUT information.
I have found that the MOST useful piece of software for manipulating images,
except NIH-Image of course, is Photoshop. If you are at a Univ you can get
really good prices on it and it happily reads almost anything I try and read
into it.

--------------------------------------


Colin it sounds like Park is at fault. We routine transfer TIFF files
from one computer Vax, Mac, PC, Sun... to another an access TIFF files
with no difficulty. My guess is that Park took short cuts and did not
properly implement TIFF (this happens alot from what I can tell).
You should demand that Park rewrite their code to properly
implement TIFF which is no at Version 6.0. T

There are lots of programs which can read TIFF files and display them
on the PC.

Nestor Z.
ANL EM Center
-------------



------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;16 Nov 1993 12:54:12 U
Received: by totalrecall.rs.itd.umich.edu (5.67/2.2)
with X.500 id AA15536; Tue, 16 Nov 93 12:54:25 -0500
Received: from anlemc.msd.anl.gov by totalrecall.rs.itd.umich.edu (5.67/2.2)
with SMTP id AA15531; Tue, 16 Nov 93 12:54:22 -0500








From: Gerald Little :      ANGJL-at-medicine.newcastle.edu.au
Date: Wed, 17 Nov 1993 08:04:58 GMT+1000
Subject: E.M. Long term storage of tissues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.931117080458.480-at-medicine.newcastle.edu.au}


"I need several options for storage of fixed tissues that might be
used later for E.M. Currently I store tissues in the original
fixative. Any other options would be appreciated."

One alternative is to put the tissue into the buffer originally used
with the fixative. The only drawback here is that after a period of
time (} 1 year) the buffer usually tends to become acidic, and there
goes the tissue preservation. Alternatively, I used this when
working years ago in another lab so have no recent knowledge of
the pros and cons, but tissue that you expect to use for TEM, process
it through osmium and up to 90% alcohol and then store. I know that
I did this and then subsequently finished processing the tissue into
epon several months later and the ultrastructural preservation was
very good. The best option to maintain good quality morphology is of
course to process the tissue into resin then you have a fossil which
can not deteriorate.

Hope this is of some help,

Gerry Little.


Dr Gerald Little | Ph (049) 215618
The Neuroscience Group |
Discipline of Anatomy | Fax (049) 216903
Faculty of Medicine |
The University of Newcastle | Email ANGJL-at-medicine.newcastle.edu.au
Australia, 2308 |




From: COOK-at-anlemc.msd.anl.gov (Russell E. Cook)
Date: Tue, 16 Nov 1993 19:19:24 -0600
Subject: SEM / Fullerenes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I passed Daniel Callahan's query to Richard Lee at Argonne. This is his
reply:

Our experience with carbon is mostly on diamond using conventional
(tungsten) filaments in an SEM. We typically resolve sub-micron filaments
and crystallites dowm to 30 to 50 nm with 10 kV operation. Fullerene
crystals were difficult for us because of charging--even at 10kV.

Our experience with an high resolution FEG-SEM was also on diamond and was
most impressive! They excel at low kV operation with no charging problems.
We have taken photos on diamond nanophase materials (coatings) with 3 nm
or better resolution.at 5kV using a new Leica-Cambridge 360FE.

You can try conventional SEM's at 5kV, but you will be limited in
magnification. If you go much above 10kV you probably will have charging
problems and loss of surface detail. The lower the kV, the better the
surface detail with a material of low Z.






From: COOK-at-anlemc.msd.anl.gov (Russell E. Cook)
Date: Tue, 16 Nov 1993 19:19:29 -0600
Subject: SEM /Fullerenes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I passed Daniel Callahan's query to Richard Lee at Argonne. This is his
reply:

Our experience with carbon is mostly on diamond using conventional
(tungsten) filaments in an SEM. We typically resolve sub-micron filaments
and crystallites dowm to 30 to 50 nm with 10 kV operation. Fullerene
crystals were difficult for us because of charging--even at 10kV.

Our experience with an high resolution FEG-SEM was also on diamond and was
most impressive! They excel at low kV operation with no charging problems.
We have taken photos on diamond nanophase materials (coatings) with 3 nm
or better resolution.at 5kV using a new Leica-Cambridge 360FE.

You can try conventional SEM's at 5kV, but you will be limited in
magnification. If you go much above 10kV you probably will have charging
problems and loss of surface detail. The lower the kV, the better the
surface detail with a material of low Z.






From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 17 Nov 1993 13:46:53 -0600 (CST)
Subject: TEM: KIKUCHI MAPS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199311171826.NAA05585-at-stc06.ctd.ornl.gov}

Reply to: RE} hex. kikuchi maps
The program DIFFRACT will do this on a Macintosh computer very nicely. This
program is available from Virtual Laboratories, 37 Highland Court., Ukiah,
Calif., USA (Attn: Janet Schlinger) FAX: 707-462-5275.

--------------------------------------

Does anyone know a programm for calculating
hexagonal kikuchi maps on PC and with/without
the possibility to vary the c/a-ratio?

Heinrich Kestler
Institut fuer Werkstoffwissenscahften
Lehrstuhl I
Martensstrasse 5
D-91058 Erlangen, Germany

Phone: 09131 857485
Fax: 09131 857504
email: kestler-at-ww.uni-erlangen.de


------------------ RFC822 Header Follows ------------------
Received: by ma160.ms.ornl.gov with SMTP;16 Nov 1993 12:48:51 -0500
Received: from anlemc.msd.anl.gov by oaunx1.ctd.ornl.gov.CTD.ORNL.GOV
(8.6.4/Ultrix3.0-C)
id MAA18295; Tue, 16 Nov 1993 12:48:47 -0500


The are also public domain programs in the EMMPDL for
calculating Diffraction Patterns/Kikuchi Maps.
They are not all elegant but they work. To access
information about the EMMPDL send an Email message to:

EMMPDL-at-anlemc.msd.anl.gov

In the Email message include the text line

SEND HELP EMMPDL

You will get instructions on how to get abstract listings
file names and how to get copies of code for the programs.

Nestor Z.
ANL EM Center
------------------------




From: ZHANG-at-bnlx1l.nsls.bnl.gov
Date: Wed, 17 Nov 1993 18:22:41 -0500 (EST)
Subject: cryo stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




USED CRYO TEM HOLDER WANTED

We are considering the use of a TEM-style cryo stage for
some x-ray
diffraction experiments on frozen hydrated specimens.
Our stability
requirements are not at all high (drift is not a problem
at the 0.1
micron/hour level), so we could probably do with a older
model cryo
stage. We would custom-build an airlock around whatever
cryo holder, so
we are not limited to a stage for e.g., JEOL or Philips.
However, we
do need a specimen loading and transfer system to go
with the
stage. We do not need much in the line of tilt control,
and copper or
other metals are fine (no need for beryllium).

If anyone has a cryo specimen holder that they are
willing to
loan or sell at modest cost, please contact me.

Thanks!
Chris Jacobsen
Department of Physics
SUNY at Stony Brook
Stony Brook NY 11794-3800
USA
fax 516-632-8101
cjacobsen-at-ccmail.sunysb.edu
CJACOBSEN-at-SBCCMAIL.bitnet






From: ridge-at-icu.ac.jp
Date: Fri, 19 Nov 1993 09:16:23 +0900
Subject: Does anyone have an old microtome?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"Microscopy Mailing List" {microscopy-at-anlemc.msd.anl.gov}

Reply_ RE} AFM: NanoIII to NIH Image And more...
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
If I label my Nanoscope III images as "test.tif" or something similar then I
can open the images into NIH-Image without losing the LUT. If I use just
"test" without the tiff extension then I get gray images. I do not have
AccessPC set to automatically open .tif files are NIH-Image files and so I am
unsure as to why this is happening. I can drag and drop a filename.tif file
onto NIH-Image and it opens just fine.
(We have also use the raw data import option but sometimes want to retain the
colors of the original image for presenation purposes.)
By the way I am getting my images off the Nanoscope by using Farallon's
Appletalk PC software, it works OK but it seems a little slow. Has anyone else
had experience with this kind of setup? I also cannot get the ethernet card I
am using (a 3COM) to work correctly with NCSA Telnet and FTP, which we would
like to use to transfer data to workstations and other computers than Macs.
--------------------------------------

Regarding John Mansfield's comment, it looks to me as though the Nanoscope III
stores raw data with a greyscale LUT. I base this on the fact that changing
color table parameters in Top View mode and then selecting another file does
not
trigger a "Current file has been modified, save it?" query the way changing the

Z scale does.

Libby Shaw
MIT CMSE Surface Analysis Facility



------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;17 Nov 1993 17:06:34 U
Received: by totalrecall.rs.itd.umich.edu (5.67/2.2)
with X.500 id AA17033; Wed, 17 Nov 93 17:06:45 -0500
Received: from anlemc.msd.anl.gov by totalrecall.rs.itd.umich.edu (5.67/2.2)
with SMTP id AA17008; Wed, 17 Nov 93 17:06:38 -0500
Message-Id: {9311172150.AA27465-at-MIT.EDU}
"Microscopy Mailing List" {microscopy-at-anlemc.msd.anl.gov}

Reply_ RE} TEM: Re: hex. kikuchi maps
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
You do have to be wary of Diffract, it is buggy, even the latest version.
There is a new product from those guys called Desktop Microscopist it is
suppoedly a complete rewrite. However, after playing with it for 10 minutes at
MSA I managed to crash it! Treat all software with care if you
havent written and debugged it yourself!!

--------------------------------------

Chris Boothroyd

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;18 Nov 1993 15:33:20 U
Received: by totalrecall.rs.itd.umich.edu (5.67/2.2)
with X.500 id AA20332; Thu, 18 Nov 93 15:33:31 -0500
Received: from anlemc.msd.anl.gov by totalrecall.rs.itd.umich.edu (5.67/2.2)
with SMTP id AA20325; Thu, 18 Nov 93 15:33:28 -0500
Message-Id: {m0p0A7u-00003QC-at-bootes.cus.cam.ac.uk}

Dear Colleagues

I have started a new teaching/research position, to find that there is a
very nice Jeol 1200, but no ultramicrotome! We have no funds for one in
the near future, and such machines in Japan are very expensive, at least
twice the price elsewhere.

However, we may be able to find enough funds for a second hand machine. I
also need a 'tome like the old JB4 for semi-thin sections, it must have a
retracting arm and be able to use glass knives. I don't care how old the
machines are as long as they work.

If anyone has an old microtome that they wish to dispose of for YEN, we
would be very interested. We will pay all shipping charges etc, of course.

Look forward to many offers!

Robert W. Ridge
Biology Department
ICU
10-2 Osawa 3-Chome
Mitaka-shi
Tokyo 181
Japan

Tel: +81-422-33-3484 (0422-33-3484 for callers within Japan)
Fax: +81-422-33-1449
Email ridge-at-icu.ac.jp






From: Louisa.Howard-at-Dartmouth.EDU (Louisa Howard)
Date: 19 Nov 93 08:15:32 EST
Subject: yeast infiltration/correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {6866666-at-blitzen.Dartmouth.EDU}

Phil:
there were two errors in the protocol I sent you. 1. You should start out
using LR White HARD grade, not medium grade; polymerize this at 50 degrees C
for 24 hrs. There may be a small of amount of fluid LR white at the top of
the capsule. These block section fine, also. 2. You may need to block free
aldehydes to prevent background problems; so you can include a step between
the buffer wash and dehydration, adding 1%ammonium chloride to your buffer
solution.
-Louisa




From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 19 Nov 1993 13:21:33 U
Subject: AFM- NanoIII

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9311191321.AA13221-at-mts-gw.pa.dec.com}
"Mikael Gustafsson" {MikGu-at-mme.liu.se}

Reply_ RE} } AFM- NanoIII to NIH Ima
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
Re: the message below that seems to be replyinmg to my gripes about slow data
transfer and printing from our Nanoscope connected to the Appletalk net at the
U of M.
Sorry, I didnt make myself clear about our physical transport layer!
We are using the Appletalk PC package on a 3-COM Ethernet card.
We have mostly Macs and several color and greyscale capable printers on the
network and so it pays us to have Appletalk connectivity.
The major problem with the networking and the Nanoscope is that the AFM
software walks over the Network software and vice versa. Digital Instruments
have fixed the AFM software so that it runs in a reduced memory mode when the
Appletalk stuff is loaded. Anyone know if I can fix this by adding more
memory? I am a PC neophyte, Macs are my computer of choice.
Plus, does anyone know why Digital Instruments used such and slow stodgy 486
for the platform for their instrument (other that it was cheap)?

OK, Opinions are mine and not the U's!
Jfm.

--------------------------------------

Hi!

Appletalk IS notoriously slow for transferring images, the maximum speed is
just over 200 Kb/sec. Compare with ethernets 10Mb/sec. Put ethernet into the
MACs... We have a fairly well working connection between PC's and UNIX
workstations (Silicon Graphics). We use Walker Richer Quinns reflection
software using TCP/IP as communication protocol. There are some very
conveniant solutions though. One includes the Lantastic network (Editirs
choice in PC-magazin for a number of years). Now there is a TCP/IP module for
this network. It supports NFS and other goodies as well. In practice, I think
this will provide more of an invisible network than ever before. Lantastic is
also available for the MAC. Moreover windows for workgroups has TCP/IP
options at very low prices.

Mikael Gustafsson
dept medical microbiology
Univ. Link0ping
MikGu-at-mme.liu.se
FAX 046/13/224789
SWEDEN

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;19 Nov 1993 06:01:36 U
Received: by hulio.liu.se (5.65c/ida.minimaster-V1.0b5d4)
id AA04183; Fri, 19 Nov 1993 11:43:21 +0100

Reply_ AFM- NanoIII
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or John_Mansfield-at-mse.engin.umich.edu
Apparently I am wrong in assuming that the computer that is used by Digital
Instruments for their Nanscope III is a cheap model. Apparently it is all
Intel built. For those of you who have an older model with a 33Mhz processor
and want a little more speed, call Mark Lean at Digital and get a price for an
Overdrive processor to take it up to 66Mhz. (or go third party).
Has anyone put a Pentium in a Nanoscope III yet? :-)






From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 19 Nov 1993 13:42:26 -0600 (CST)
Subject: Images: TN 8500

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Pat Robinson asks......
....................stuff deleted.............

} this 8500 is quite a different beastie. I'm especially interested in
} ways and means of moving images to our network of Digital (DEC) PC's
} running desktop publishing software, via TIFF or otherwise.

If you really want to work efficiently. Toss the TN buy yourself
a Mac and framegrabber board. Then get a copy of the Public Domain
program NIH-Image. It will grab images off your Sony, do some
reasonable processing and save files as Tiff which you can then send
via your network (use Pathworks, AppleTalk, TCP, just chooze your
protocol & hardware) to your DEC machines.

Nestor Z.
ANL EM Center






From: jacobsen-at-xray1.physics.sunysb.edu (Chris Jacobsen)
Date: Fri, 19 Nov 1993 16:21:13 -0500 (EST)
Subject: Image Printing (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In reference to
} From jkelly-at-pruffle.nist.gov Fri Nov 19 15:58:12 1993
} To: microscopy-at-anlemc.msd.anl.gov
} Subject: Image Printing
}
} I am working with image files on PC's - usually TIFF format
} on IBM compatibles. I would like to get good quality B & W
} plain paper hardcopy of the images. What are the good
} combinations of print software and printers for either the PC's
} or Mac environment? Thanks in advance.

For computer image printing, two popular solutions are laser printers
(dithered black and white) and dye sublimation printers. We use IDL
from Research Systems Inc. (Boulder CO) to do many many things, one of
which is to generate PostScript files of images. It runs on PC and
Mac (and Unix and VMS). We then get pretty good and very cheap B&W
prints on any old PostScript laser printer. For the final print, we
have a Tektronix Phaser IIsd (the "sd" or "sdx" is important) dye
sublimation printer which takes color or B&W PostScript files. Kodak
also now sells a similar product. Codonics sells a dye sublimation
printer which takes TIFF files directly.


--
*******************
Chris Jacobsen, Asst. Prof., Department of Physics, SUNY at Stony Brook
Phone (516) 632-8093, FAX -8101 Bitnet: cjacobsen-at-sbccmail
jacobsen-at-xray1.physics.sunysb.edu ALL-IN_ONE: CJACOBSEN
*******************




From: Bernhardt Sainieidukat :      sainieid-at-badlands.NoDak.edu
Date: Fri, 19 Nov 1993 16:33:27 -600 (CST)
Subject: SEM sample prep of powders

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9311192200.AA20344-at-mts-gw.pa.dec.com}

I have some synthetically prepared mineral powders, grain size
in the 1-30 micron range, which I would like to get some EDS analyses of.
This would mean embedding the microcrystals in epoxy and polishing them to
get nice cross sections -- would someone with experience in this kind of
sample prep be willing to share some tips/hints?

Thanks,

Bernhardt Saini-Eidukat
Dept. of Geosciences
North Dakota State University
Fargo, ND 58105







From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 19 Nov 1993 20:15:30 -0600 (CST)
Subject: Image Output

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Adding to the discussion of printing we use the following
for output.


Sony Vidoeo Printer - direct from TV monitors - Poor quality
Notebook records (4x 5) output

Apple Laserwriter II NTX - in photograde mode - Reasonable B&W
dithered. Used a prescreen computer output on the
following.

Tek Phaser II SDX - High quality- Dye Sub B&W , & Color
excellent positives & viewgraphs
used for output from computers

Lasergraphics Film Writer - 35mm film & slides, Polaroid, & 4 x 5 format
used for computer output


Images are either directly digitized by
serial scanning - On-line TEM/SEM
slow scan CCD - On-line TEM
TV rate CCD - On-line TEM, & Off-line Photostand
Conventional Vidicon - ON-line TEM & Off-line Photostand
Flatbed Scanners - Off-line
Microtek 300G- Prints,
XRS/Microtek 300 GX - Prints & Negatives

Nestor Z.
ANL EM Center






From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 19 Nov 1993 20:22:46 -0600 (CST)
Subject: TEM: Philips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Don't know of any free Philips scopes. Have your
called Philips Service Dept? Their number is 201-529-3800.

They sometimes have a line on instruments being given away.





From: Charles E. Lyman :      cel1-at-Lehigh.EDU
Date: Mon, 22 Nov 1993 15:29:44 -0500
Subject: Reply to: Image Analysis T/N 8500 system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Patrick Robinson,

We have been transferring files from the Tracor 8500 to PC and Mac formats using
the following steps:

1. IBM-compatible computer connected from serial port 7 of 8500 to com2 of IBM.
2. On 8500 convert file to ".tif" format and re-save the file to the EXPORT
directory.
3. Files are transferred to an IBM-compatible image handling application, such
as Procomm. If using Procomm, open Procomm and input the following file
specification: "/u3/export/name.tif". As the file is transferred, a byte count
is tallied which should match that read on the 8500.
4. Copy image from Procomm directory to an IBM-formatted diskette.

Cheers, Charles E. Lyman


Department of Materials Science and Engineering
Lehigh University, 5 East Packer Avenue
Bethlehem, PA 18015-3195
E-mail: cel1-at-Lehigh.edu Tel: 215-758-4249 FAX: 215-758-4244





From: Glenn Poirier :      GLENN_P-at-geosci.lan.mcgill.ca
Date: Mon, 22 Nov 1993 10:54:06 EST5EDT
Subject: Re: SEM sample prep of powders

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9311221558.AA20274-at-sifon.cc.mcgill.ca}
To: Bernhardt Sainieidukat {sainieid-at-badlands.NoDak.edu}

Bernhardt,
I've mounted small grains for use in the microprobe on several
occasions. Usually the material I work with is a bit larger, but the
techniques should still apply if you work carefully.
If you have a large sample and are not concerned with individual
grains, the simplest way to work is to set them in in epoxy on a
standard thin section blank. If you have a reliable thin-section
technician he should have no problen grinding it down to a 15 um
thickness.
Alternatively, if you have a small number of grains or you are
interested in specific grains you can mount them in a plexiglass
holder. First cut out a piece of plexiglass to fit into your sample
holder (the thinner the better, I find 3 mm is about right). Next drill a
number of small holes in the holder (I use a 2mm diameter). I mount
these holders on a cleaned thin section blank using double sided
tape. You should be very careful that there are no bubbles between
the tape and the blank. Fill a hole with epoxy (I use Struers Epofix,
it's nice and thin). You probably will want to carry out the next
steps under a microscope. Using very fine forceps drop a grain into a
hole and stir the epxoy with a fine wire to wet the grain and cause it to
settle into the epoxy. If necessary you can adjust the grain so that it's
resting on its most stable surface. Depending on the grain size I can
usually get 10-15 grains in a mount before they start interfering with
each other. Once the epoxy has set you can remove the blank slide
on the bottom (usually I have to break it off).
Polishing these mounts is crucial, if you take to much material
off you lose your sample (This can really ruin your day if that was the
only sample you had). I generally start with a very light brush with
300 grit SiC paper, just enough to remove any excess epoxy. This is
followed by 600 grit paper. At this stage I polish for 30 seconds to a
minute and then look at the result in a microscope to see how much
exposed material I have. When sufficient material is exposed I do
the fine polishing using 3, 1 and then 0.3 um Al2O3 either by hand or
on a lap, depending how nervous I am about losing the material.
Again I use short polishing periods followed by observation in the
microscope. I've generally had good results with this technique and
have rarely lost any material. One problem I have with larger grains
is that it is sometimes difficult to get a large polished surface when
the grain is sitting on it edges.
I hope this helps you. Call me back if any of this is not clear

**********************************************************************
* Glenn Poirier glenn_p-at-geosci.lan.mcgill.ca *
* Electron Microprobe Lab Phone: (514) 398 6774 *
* Earth and Planetary Sciences Fax: (514) 398 4680 *
* McGill University THERE ARE THREE SIDES TO EVERY STORY; *
* Montreal, Quebec YOUR SIDE MY SIDE AND THE TRUTH *
**********************************************************************




From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Tue, 23 Nov 1993 08:48:07 +0000 (GMT)
Subject: LM - Ultra long working distance objectives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

I'm looking to buy a set of ultra long working distance objectives to fit
onto an Olympus BHSM microscope. The longest WDs I've found so far belong
to Olympus' own objectives (20x with 11mm and 50x with 8mm). Anyone know of
any alternatives?

Thankyou,

Keith Hallam






From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Tue, 23 Nov 1993 08:48:07 +0000 (GMT)
Subject: LM - Ultra long working distance objectives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {28025.9311230848-at-irix.bris.ac.uk}

Dear All,

I'm looking to buy a set of ultra long working distance objectives to fit
onto an Olympus BHSM microscope. The longest WDs I've found so far belong
to Olympus' own objectives (20x with 11mm and 50x with 8mm). Anyone know of
any alternatives?

Thankyou,

Keith Hallam






From: Jeff Sweeney :      sweeney-at-dionheinz.uchicago.edu
Date: Tue, 23 Nov 1993 08:29:57 -0600
Subject: LM - Ultra long working distance objectives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} I'm looking to buy a set of ultra long working distance objectives to
} fit onto an Olympus BHSM microscope. The longest WDs I've found so far
} belong to Olympus' own objectives (20x with 11mm and 50x with 8mm).
} Anyone know of any alternatives?

One trick is to look for U-stage lenses. These are designed for
universal-stage mineralogical microscopes where the stage tilts as well
as rotates. These lenses have a separate hemispherical glass element
that is placed over the thin-section. Used without the hemisphere they
have very long working distances.

Another trick is to use reflecting objectives.

I don't know the Olympus BHSM. Assuming it has standard threads (.8in
38tpi) and the equivalent of a 160mm tube length... We use a Leitz UT40
which is a U-stage lens. Without the hemisphere it has a magnification
of 25x and a working distance of 14mm. Leitz has a few other U-stage
lenses that are parfocal with this lens. Chromatic aberration can be a
problem for refracting objectives with these long working distances.

We also use a reflecting objective from Ealing (15x, 21mm or 24mm
working distance). Ealing sells several other reflecting objectives
with higher magnification. The 15x is identical to the 15x lens made
by Beck as far as I can tell. You may be able to purchase the Ealing
lens corrected for infinity focus or a 210mm tube length, if that is
what the Olympus requires. These reflecting objectives are just the
ticket if you can tolerate their large size and the central obscuration.

Jeff Sweeney sweeney-at-dionheinz.uchicago.edu

Some good references are:

%0 Journal Article
%A Burch, C. R.
%D 1943
%T On aspheric anastigmatic systems
%B Proceedings of the Physical Society
%V 55
%N 312
%P 433-444
%K reflecting objective


%0 Journal Article
%A Burch, C. R.
%D 1943
%T Reflecting microscopes
%B Proceedings of the Physical Society
%V 59
%P 41-46
%K reflecting objective


%0 Journal Article
%A Burch, C. R.
%D 1945
%T Flat-field singlet aplanats
%B Proceedings of the Physical Society
%V 57
%P 567-576
%K reflecting objective


%0 Journal Article
%A Burch, C. R.
%D 1947
%T Semi-aplanat reflecting microscopes
%B Proceedings of the Physical Society
%V 57
%P 47-49
%K reflecting objective




From: NAME \ Greg Erdos, ICBR EM Core Lab. :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Tue, 23 Nov 1993 11:38:45 -0500 (EST)
Subject: Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Position available for a Biological Electron Microscopy Technician. Preference given to one with experience with plant tissue

Location: Lake Alfred Florida

Contact: Personnel Office
Department of Citrus
P.O. Box 148
Lakeland, FL 33802
(813) 499-2476

` DEADLINE: Dec 10, 1993

Pay Grade 18 ($20,553 to $33,893)




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 23 Nov 1993 10:54:17 -0600 (CST)
Subject: TEM: Electropolishing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Stefan Straub asked about electropolishing of Ti alloys
and 12%Cr Steel

Bernie Kestel ANL's expert at electropolishing suggest the
following:

Ti alloys:

13% HCL/87%Methanol -at- -50 C, 90 Volts/35 mA in a
single jet electropolishing machine

or

60ml Percholoric Acid
590 ml Methanol
350 ml 2-butoxy ethanol (butyl cellosolve)
-at- 0 C, 40-50 Volts/50 mA in a
single jet electropolishing machine

12%Cr.Steel Alloy

look up Ultramicroscopy Vol 26, (1988) pge 405-408

by Kestel & Wiggins

60 ml perchloric acid
460 ml ethyl alcohol
280 ml n-butyl alcohol
100 ml butyl cellosolve
-15 C , 100 V

Nestor Z (for Bernie Kestel)
ANL EM Center
==============================================================




From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Tue, 23 Nov 1993 18:02:38 +0000 (GMT)
Subject: Re: LM - Ultra long working distance objectives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {22347.9311231802-at-irix.bris.ac.uk}

A couple of things I forgot to mention in my original mailing - the Olympus has
infinity corrected optics (I had found that Nikon had a nifty lens, but it
turned out not to be infinity corrected) and the microscope is part of a
Renishaw imaging Raman spectrometer. I have been told by others that
reflecting objectives would be no use since the Raman laser would either be
reflected straight back into the spectrometer without reaching my sample or
would have to be defocussed, preventing me doing any small area analysis.

Keith







From: NAME \ Greg Erdos, ICBR EM Core Lab. :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Tue, 23 Nov 1993 15:17:30 -0500 (EST)
Subject: Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Position available for a Biological Electron Microscopy Technician. Preference given to one with experience with plant tissue

Location: Lake Alfred Florida

Contact: Personnel Office
Department of Citrus
P.O. Box 148
Lakeland, FL 33802
(813) 499-2476

` DEADLINE: Dec 10, 1993

Pay Grade 18 ($20,553 to $33,893)




From: COOK-at-anlemc.msd.anl.gov
Date: Tue, 23 Nov 1993 14:33:33 -0600 (CST)
Subject: SEM: Fullerenes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Richard Lee of the Energy Technology Division at Argonne National Laboratory
answers Daniel Callahan's question of 11 Nov 93:

Our experience with carbon is mostly on diamond with tungsten filaments in a
SEM. We typically resolve sub-micron filaments and crystallites down to 30 to
50 nm at 10kV. Fullerene crystal were difficult for us because of charging,
even at 10kV.

Our experience with an high resolution FEG-SEM was with diamond and the results
were impressive. The FEG-SEM excels at low kV with no charging problems. We
have taken photos of diamond nanophase materials (coatings) with 3 nm or better
resolution at 5 kV using a new Leica-Cambridge 360FE.

You can try conventional SEM's at 5 kV but you will be limited in magnification.
If you go much above 10 kV you probably will have charging problems and loss of
surface detail. The lower the kV the better the surface detail with a material
of low Z.




From: Jeff Sweeney :      sweeney-at-dionheinz.uchicago.edu
Date: Tue, 23 Nov 1993 15:14:14 -0600
Subject: LM - Ultra long working distance objectives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} A couple of things I forgot to mention in my original mailing - the
} Olympus has infinity corrected optics (I had found that Nikon had a
} nifty lens, but it turned out not to be infinity corrected) and the
} microscope is part of a Renishaw imaging Raman spectrometer. I have
} been told by others that reflecting objectives would be no use since
} the Raman laser would either be reflected straight back into the
} spectrometer without reaching my sample or would have to be
} defocussed, preventing me doing any small area analysis.

Geez, you're making this more difficult :-) Chromatic aberration is not
a problem with Ramam since you are working so near the Rayleigh peak, so
look for a U-stage lens with infinity focus. You can use a reflecting
objective but you must send the probe laser into the lens off-axis.
Your imaging system may not accomodate this (probably not, huh). The
focus of the laser can be adjusted with the beam expander if your beam
expander is adjustable (probable not, huh). Alternatively, the laser
can be prefocused with a long focal-length plano-convex lens. You might
consider coating the reflecting objective for the laser wavelength.
This would help aviod beam damage (not really a problem) and would give
better signal in general.

Jeff Sweeney sweeney-at-dionheinz.uchicago.edu




From: e-reuter-at-uiuc.edu (Erik Reuter)
Date: Tue, 23 Nov 1993 19:06:38 -0500
Subject: LM: Need basic texts/review articles on design

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199311240105.AA13285-at-ux1.cso.uiuc.edu}
X-Sender: reuter-at-ux1.cso.uiuc.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


I would like to hear recommendations for text books and review articles on
the subject of light microscopy system design. I am interested in both
basic information and the most recent technology. I would specifically be
interested in hearing about those "desk reference" type books or review
articles you always keep near by, and articles that have good
bibliographies.

I am working on designing a photoluminescence imaging system (not confocal).

Please send references to me and I will compile a list. Also, if anyone
would like a copy of the completed list, just let me know. Thanks.


Erik Reuter, e-reuter-at-uiuc.edu, 217-244-6378 phone, 217-244-6375 fax






From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Wed, 24 Nov 1993 16:45:21 +0000 (GMT)
Subject: Re: LM - Ultra long working distance objectives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Geez, you're making this more difficult :-)i

And I haven't yet mentioned the high temperatures I want to have my samples
at! Also, one of our third year students wants to Raman his diamond
coatings at 400 C - What will that do to either ULWD or standard objectives???

Chromatic aberration is not
} a problem with Ramam since you are working so near the Rayleigh peak, so
} look for a U-stage lens with infinity focus. You can use a reflecting
} objective but you must send the probe laser into the lens off-axis.
} Your imaging system may not accomodate this (probably not, huh).

I'll go and ask.

The
} focus of the laser can be adjusted with the beam expander if your beam
} expander is adjustable (probable not, huh).

I already know this is, but you might be able to tell we haven't had the
system long and are all still coming to terms with it. Someone here
suggested using reflecting objectives with the beam expanded and with
a blanking-widget in the lens to prevent direct reflection of the laser back into the spectrometer.

Alternatively, the laser
} can be prefocused with a long focal-length plano-convex lens. You might
} consider coating the reflecting objective for the laser wavelength.
} This would help aviod beam damage (not really a problem) and would give
} better signal in general.
}
} Jeff Sweeney sweeney-at-dionheinz.uchicago.edu
}
iThankyou,

Keith






From: COOK-at-anlemc.msd.anl.gov (Russell E. Cook)
Date: Wed, 24 Nov 1993 10:59:10 -0600
Subject: LM: Basic texts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Erik, check out a couple of Kodak publications:

"Photography Through The Microscope", P-2, cat.# 152-8371
"Kodak Scientific Imaging Products", L-10, cat. # 813-9321

The first has loads of basic information about objectives, condensers,
setting-up illumination, flourescence, phase contrast, etc. Both have
bibliographies.






From: colijn-at-kcgl1.eng.ohio-state.edu (Henk Colijn)
Date: Wed, 24 Nov 1993 16:54:22 -0400 (EDT)
Subject: Available: JEOL 200CX TEM/STEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Available: JEOL 200CX TEM/STEM

The scope is 10 years old but in very good condition. It has been
continuously maintained under service contract and is still in daily use.
The components and accessories are listed below. All documentation and
records are included. We would like to sell this as a complete package.

Contact:
Hamish Fraser,
(614)292-2708, email: fraser.3-at-osu.edu
Henk Colijn
(614)292-0674, email: colijn.1-at-osu.edu


The equipment included with the microscope is

JEM-200CX TEM - S/N EM132031-86
200CX-SEGZ JEOL 200CX side entry goniometer
HM-PP 3.5A, +30o SEG polepiece
HA-PP 4.5A, +60o SEG polepiece
200CX-SQH-2 High Mag single tilt holder
200CX-SCSH Std. SEG quick change single tilt holer
200CX-BST double tilt holder (w/ one Gatan Be cup)
200CX-SHH sample heating holder
200CX-SCH single tilt cooling holder
200CX-SHU heating holder control unit
200CX-SEH single tilt straining holder
200CX-SFH faraday cup
Haskris recirculating water chiller
200CX-ASID-3D STEM/SEM system
200CX -BEI-3 BSE detector (needs new Si crystals)
200CX-BF/DF STEM BF/DF option
200CX-FLC Free lens control
35-GMC Gamma control
HA-EDS high angle EDS interface
200CX-HXS hard x-ray aperture
200CX-IMS-2 Image selector switch
35-MDD Multiple image display
200CX-SRT Scan Rotation & tilt correction
200CX-UHR UHR (photo) CRT w/ Polaroid 545 film back
35-VCA Video control amp. (derivative/filtering)
200CX-WFM-B Tek 501 waveform monitor
200CX-WFM-M lens current readout
35-YMD Y modulation device
EM-THG Top entry Ultrahigh resolution goniometer
TN-2000 Tracor-Northern TN-2000 EDS system (PDP-11/23 w/ 128kB RAM)
SpectraChrome 512 Color monitor
Horizontal Be window EDS detector
High Takeoff angle (72o) Be window EDS detector
EDS preamplifiers (x2)
Digital beam control and interface
EDS and Image acquisition software
DEC LA-120 printer
Honeywell Video Graphics recorder
JEOL ASD system (projector lens scan coil) with OSU mods
for scanning the TEM image)
OSU built hollow cone illumination device
Manuals and notes
misc spare parts and filaments, including pole pieces and
defl. coils
HP 7221C 8 pen plotter
operator's chair
Gatan DT cryo low bkg holder
Gatan cryo transfer holder






colijn.1-at-osu.edu OSU Campus Electron Optics Facility 292-0674
-------------------------------------------------------------------
Assumption is the mother of all screwups.





From: SMITHC-at-ANIMAL.SSEC.HONEYWELL.COM
Date: Wed, 24 Nov 1993 16:12:40 -0600 (CST)
Subject: re: Image Printing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Regarding J. Kelly's (jkelly-at-enh.nist.gov) question:

} I am working with image files on PC's - usually TIFF format
} on IBM compatibles. I would like to get good quality B & W
} plain paper hardcopy of the images. What are the good
} combinations of print software and printers for either the PC's
} or Mac environment? Thanks in advance.

Have you considered videoprinters? You can videoprint any
analog signal: RS-170 (NTSC), PAL (Europe), VGA... at 1200
dots/inch, 64 grey-levels, and it's *fast* (30 sec) compared
to digital printing. You get near Polaroid quality for 40
cents a shot (one 8x10 inch, or two different 3x5 inch images
can be printed together). I know you specified "plain paper,"
but the thermal paper is easily handled. Our
Seikosha VP-3500 has been so reliable and popular
with our customers, we put another on a Sun workstation based SEM.
Capital investment is about $7500. One drawback: the images
are not as stable. When I put a cup of hot coffee down on
a print, it "developed" a black circle.
*******************
Craig A. Smith, Honeywell Solid State Electronics Center, MN14-2C25
12001 State Highway 55, Plymouth, MN 55441-4799 USA
Phone (612) 954-2895, FAX (612) 954-2040
smithc-at-ccsvax.ssec.honeywell.com
*******************






From: Nancy L. Desmond :      nld-at-galen.med.virginia.edu
Date: Fri, 26 Nov 1993 14:16:39 -0500
Subject: gray scale printers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Like J. Kelly, we are interested in printers. Specifically we
are interested in purchasing a high-resolution gray scale printer
to print TIF files from both PC's and from Silicon Graphic
computers. Our files contain light microscopic images of
hippocampal dendrites visualized at high magnification using a
variety of staining methods (Golgi, biocytin, neurobiotin). I
have information from Harris on their PhotoPro2000 and from Alden
on their 9315 continuous tone printer. They range in price from
$7500-10,000. What has been your experience & what would you
recommend? We need a high-res image, and I am a bit concerned
about the longevity of the thermal paper images. Many thanks.

Nancy L Desmond
Dept. of Neurosurgery
Univ. of Virginia Health Sciences Center
Charlottesville, VA 22908
804.924.5607 (phone)
NLD-at-GALEN.MED.VIRGINIA.EDU




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sat, 27 Nov 1993 11:10:14 -0600 (CST)
Subject: Images:Printing Thermal Paper

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nancy Desmonds writes....

....deleted text.......

} I have information from Harris on their PhotoPro2000 and from Alden
} on their 9315 continuous tone printer. They range in price from
} $7500-10,000.

This is the first I've hear of these brands. So I have
no opinion, except that they sound expense for thermal
printers.
Lots of systems were exhibited at the MSA annual meeting
for continuous tone printing. You may
want to get a copy of the Exhibitors Guide from the
MSA meeting and call up a few of the vendors for more
information. Try contacting Bill Gunning
Dept. of Path. Medical College of Ohio. 3000 Arlington
Ave. Toledo, Ohio 43699-0008. He runs the Advertising
for the MSA EXPO and can probably tell you how to
get a copy. (419) 381-3484

} We need a high-res image, and I am a bit concerned
} about the longevity of the thermal paper images.

High res means different things to different people.
Do you mean pixel resolution, gray scale resolution,
Image size or what? It also depends upon you digitization
source (TV, CCD, Scanner....) and it's "pixel/gray scale" resolution.
Also what is the purpose of your prints? Publication
reports, records, or what.. If you have them on the
computer it's certainly archived there resonably well
at least as good as the lifetime of data on disks. I've
got floppy disks with data on them that are now approaching
15 & 20 years old and to my surprize the data was still readable
even though the disks at the time were only supposed to
be good for ~ 10 years! Of course that implies you still
have the computer to read all that "good" old data :-)

I would not trust thermal paper for archiving but only
for a quick look see and maybe a lab notebook record.
The best thing to do here is get the manufacturer's
your interested in to give you copies of their prints
and lay them out in the sunlight and shade for a few days/weeks.
See how long they survive and then decide. The different
brands will give different results, some are suprizingly
good.


Nestor Z.
ANL EM Center






From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sat, 27 Nov 1993 11:51:16 -0600 (CST)
Subject: TEM:Celli Tape

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Re: TEM specimen preparation of Graphite:

Here's a question for those of you that used to use the
old Celli tape popular in England about 10+ years ago.

In the past I had some of this tape which I used to use
to prepare specimens of graphite for TEM. Here I would
use the old trick of just continually touching the surface
of the graphite with fresh tape to delaminate the graphite
gradually until I made a nice thin & optically transparent
section, which stuck to the tape and was usually perfect for
TEM. The nice thing about this old type of scotch tape was that
the adhesive used would disolve beautifully in Acetone and the
backing for that tape would just float away as it was not
soluablein the acetone. Unfortuantely, the current type
of "SCOTCH BRAND" transparent tape completely dissolves
in Acetone and most other solvents I tried.

I'm back to trying to make more graphite samples as
my old ones have finally bit the dust so to speak.
Since the new scotch tape failed, I tried extraction
replica tape which is available from most EM supply
vendors, however it does not have the adhearance
of either the old Celli tape or the newer scotch tapes.
(BTW the new scotch tape very nicely delaminated the graphite
and gives beautifully thin sections if I could ever
remove them from the adhesive). Does anyone know of a
solvent which works on the newer scotch brand tapes
or have another/similar idea/procedure for layered
structures which are not strongly bound?.

I'd prefer to avoid ion milling the graphite although
I may have to resort to this in the end.

Nestor Z.
ANL EM Center





From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Mon, 29 Nov 1993 12:21:41 +0000 (GMT)
Subject: LM - ULWD objectives summary

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9836.9311291222-at-irix.bris.ac.uk}

Forwarded message:
From K.R.Hallam-at-bristol.ac.uk Mon Nov 29 12:21:47 1993
Message-Id: {9815.9311291221-at-irix.bris.ac.uk}

Dear All,

I was asked if I would post a summary of the replies received after my
queation over ULWD objectives to attach to an Olympus BHSM / Renishaw
imaging Raman system. Here goes....


} From jacobsen-at-xray1.physics.sunysb.edu Tue Nov 23 14:22:40 1993

Nikon makes a toolmaker's lens which can be fitted onto
a Nikon Optiphot. It has 10x mag and a working distance
of several cm. Very good images...


} From sweeney-at-dionheinz.uchicago.edu Tue Nov 23 15:26:16 1993

One trick is to look for U-stage lenses. These are designed for
universal-stage mineralogical microscopes where the stage tilts as well
as rotates. These lenses have a separate hemispherical glass element
that is placed over the thin-section. Used without the hemisphere they
have very long working distances.

Another trick is to use reflecting objectives.

I don't know the Olympus BHSM. Assuming it has standard threads (.8in
38tpi) and the equivalent of a 160mm tube length... We use a Leitz UT40
which is a U-stage lens. Without the hemisphere it has a magnification
of 25x and a working distance of 14mm. Leitz has a few other U-stage
lenses that are parfocal with this lens. Chromatic aberration can be a
problem for refracting objectives with these long working distances.

We also use a reflecting objective from Ealing (15x, 21mm or 24mm
working distance). Ealing sells several other reflecting objectives
with higher magnification. The 15x is identical to the 15x lens made
by Beck as far as I can tell. You may be able to purchase the Ealing
lens corrected for infinity focus or a 210mm tube length, if that is
what the Olympus requires. These reflecting objectives are just the
ticket if you can tolerate their large size and the central obscuration.

Some good references are:

%0 Journal Article
%A Burch, C. R.
%D 1943
%T On aspheric anastigmatic systems
%B Proceedings of the Physical Society
%V 55
%N 312
%P 433-444
%K reflecting objective

%0 Journal Article
%A Burch, C. R.
%D 1943
%T Reflecting microscopes
%B Proceedings of the Physical Society
%V 59
%P 41-46
%K reflecting objective

%0 Journal Article
%A Burch, C. R.
%D 1945
%T Flat-field singlet aplanats
%B Proceedings of the Physical Society
%V 57
%P 567-576
%K reflecting objective

%0 Journal Article
%A Burch, C. R.
%D 1947
%T Semi-aplanat reflecting microscopes
%B Proceedings of the Physical Society
%V 57
%P 47-49
%K reflecting objective


} From treado+-at-pitt.edu Tue Nov 23 21:03:41 1993

In response to your question posted to the Microscopy discussion list
November 23, we use Leica ultra long working distance infinity-corrected
objectives on our Olympus BHS microscope.


} From sweeney-at-dionheinz.uchicago.edu Tue Nov 23 21:30:36 1993

Geez, you're making this more difficult :-) Chromatic aberration is not
a problem with Ramam since you are working so near the Rayleigh peak, so
look for a U-stage lens with infinity focus. You can use a reflecting
objective but you must send the probe laser into the lens off-axis.
Your imaging system may not accomodate this (probably not, huh). The
focus of the laser can be adjusted with the beam expander if your beam
expander is adjustable (probable not, huh). Alternatively, the laser
can be prefocused with a long focal-length plano-convex lens. You might
consider coating the reflecting objective for the laser wavelength.
This would help aviod beam damage (not really a problem) and would give
better signal in general.


} From e-reuter-at-uiuc.edu Wed Nov 24 00:20:00 1993

I am also interested in learning about ultra long working distance
objective. Would you please summarize to the list or to me what you find
out in a few days?


} From timonf-at-earth.ruu.nl Wed Nov 24 07:41:09 1993

The problem with U-stage lenses is that the highest magnification I know
which is available is a 30 x, and the optical quality is not too good (some
distortion of the image and loosing a loss of brightness)


} From muepf-at-iff067.iff.kfa-juelich.de Wed Nov 24 07:52:31 1993

One of the --- expensive --- alternatives is to get a Questar long

[Sorry - I've lost part of the original here because someone picked up
the telephone next door while my modem was connected - the message went on
to mention a 3000 ECU alternative - Could you please resend the details to
me - Thank you]


} From treado+-at-pitt.edu Wed Nov 24 17:14:22 1993

If you are only interested in doing Raman microprobing the reflective
objective may be suitable by defocusing and illuminating off-axis. If you
want to perform imaging I would use refractive objectives. The imaging
quality is far superior with refractive optics.

We use Leica objectives: 5X, 0.10 NA, 50 mm WD and 50X, 0.45 NA, 20.6 mm WD.

We operate at elevated temperatures (} 200 oC) and jacket the objective in
a water cooled housing. According to Leica the optical cements fail above
110 degrees.


Keith Hallam







From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Mon, 29 Nov 1993 12:21:41 +0000 (GMT)
Subject: LM - ULWD objectives summary

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-anlemc.msd.anl.gov

Forwarded message:
From K.R.Hallam-at-bristol.ac.uk Mon Nov 29 12:21:47 1993
Message-Id: {9815.9311291221-at-irix.bris.ac.uk}

Dear All,

I was asked if I would post a summary of the replies received after my
queation over ULWD objectives to attach to an Olympus BHSM / Renishaw
imaging Raman system. Here goes....


} From jacobsen-at-xray1.physics.sunysb.edu Tue Nov 23 14:22:40 1993

Nikon makes a toolmaker's lens which can be fitted onto
a Nikon Optiphot. It has 10x mag and a working distance
of several cm. Very good images...


} From sweeney-at-dionheinz.uchicago.edu Tue Nov 23 15:26:16 1993

One trick is to look for U-stage lenses. These are designed for
universal-stage mineralogical microscopes where the stage tilts as well
as rotates. These lenses have a separate hemispherical glass element
that is placed over the thin-section. Used without the hemisphere they
have very long working distances.

Another trick is to use reflecting objectives.

I don't know the Olympus BHSM. Assuming it has standard threads (.8in
38tpi) and the equivalent of a 160mm tube length... We use a Leitz UT40
which is a U-stage lens. Without the hemisphere it has a magnification
of 25x and a working distance of 14mm. Leitz has a few other U-stage
lenses that are parfocal with this lens. Chromatic aberration can be a
problem for refracting objectives with these long working distances.

We also use a reflecting objective from Ealing (15x, 21mm or 24mm
working distance). Ealing sells several other reflecting objectives
with higher magnification. The 15x is identical to the 15x lens made
by Beck as far as I can tell. You may be able to purchase the Ealing
lens corrected for infinity focus or a 210mm tube length, if that is
what the Olympus requires. These reflecting objectives are just the
ticket if you can tolerate their large size and the central obscuration.

Some good references are:

%0 Journal Article
%A Burch, C. R.
%D 1943
%T On aspheric anastigmatic systems
%B Proceedings of the Physical Society
%V 55
%N 312
%P 433-444
%K reflecting objective

%0 Journal Article
%A Burch, C. R.
%D 1943
%T Reflecting microscopes
%B Proceedings of the Physical Society
%V 59
%P 41-46
%K reflecting objective

%0 Journal Article
%A Burch, C. R.
%D 1945
%T Flat-field singlet aplanats
%B Proceedings of the Physical Society
%V 57
%P 567-576
%K reflecting objective

%0 Journal Article
%A Burch, C. R.
%D 1947
%T Semi-aplanat reflecting microscopes
%B Proceedings of the Physical Society
%V 57
%P 47-49
%K reflecting objective


} From treado+-at-pitt.edu Tue Nov 23 21:03:41 1993

In response to your question posted to the Microscopy discussion list
November 23, we use Leica ultra long working distance infinity-corrected
objectives on our Olympus BHS microscope.


} From sweeney-at-dionheinz.uchicago.edu Tue Nov 23 21:30:36 1993

Geez, you're making this more difficult :-) Chromatic aberration is not
a problem with Ramam since you are working so near the Rayleigh peak, so
look for a U-stage lens with infinity focus. You can use a reflecting
objective but you must send the probe laser into the lens off-axis.
Your imaging system may not accomodate this (probably not, huh). The
focus of the laser can be adjusted with the beam expander if your beam
expander is adjustable (probable not, huh). Alternatively, the laser
can be prefocused with a long focal-length plano-convex lens. You might
consider coating the reflecting objective for the laser wavelength.
This would help aviod beam damage (not really a problem) and would give
better signal in general.


} From e-reuter-at-uiuc.edu Wed Nov 24 00:20:00 1993

I am also interested in learning about ultra long working distance
objective. Would you please summarize to the list or to me what you find
out in a few days?


} From timonf-at-earth.ruu.nl Wed Nov 24 07:41:09 1993

The problem with U-stage lenses is that the highest magnification I know
which is available is a 30 x, and the optical quality is not too good (some
distortion of the image and loosing a loss of brightness)


} From muepf-at-iff067.iff.kfa-juelich.de Wed Nov 24 07:52:31 1993

One of the --- expensive --- alternatives is to get a Questar long

[Sorry - I've lost part of the original here because someone picked up
the telephone next door while my modem was connected - the message went on
to mention a 3000 ECU alternative - Could you please resend the details to
me - Thank you]


} From treado+-at-pitt.edu Wed Nov 24 17:14:22 1993

If you are only interested in doing Raman microprobing the reflective
objective may be suitable by defocusing and illuminating off-axis. If you
want to perform imaging I would use refractive objectives. The imaging
quality is far superior with refractive optics.

We use Leica objectives: 5X, 0.10 NA, 50 mm WD and 50X, 0.45 NA, 20.6 mm WD.

We operate at elevated temperatures (} 200 oC) and jacket the objective in
a water cooled housing. According to Leica the optical cements fail above
110 degrees.


Keith Hallam







From: PO-at-, parmly1.parmly.luc.edu-at-parmly1.parmly.luc.edu
Date: 29 Nov 93 10:44:59 CST6CDT
Subject: Re: Cameras & devices for higher frame-rates than standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.931129104500.288-at-parmly1.parmly.luc.edu}
To: Microscopy-at-anlemc.msd.anl.gov


} We are doing microscopic image processing to access physical properties
} (and their change) of lipid membranes and polymers in different setups.
} In the near future we will face the problem to escape the limits of standard
} video frequency (30/25 Hz). Therefore I want to ask whether anybody is
} listening, who has experience with higher frame-rate cameras and other
} devices and who would be willing to give me some advice in this field.
Question deleted...
} I am quite unsure whether the answers to this query might be of general
} interest or not, so please feel free to email me directly at
} mschindl-at-physik.tu-muenchen.de.
I think this is of interest...I would certainly like to know the
answers to these questions.
Phil Oshel




From: naresh-at-funky.mm.uky.edu
Date: Mon, 29 Nov 1993 08:13:01 -0500
Subject: Re: TEM:Celli Tape

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





---- Transcript of session follows ----
%UCX-E-SMTP_NOSUCHNODE, Remote host unknown, anlemc.msd.anl.gov
-SYSTEM-F-TIMEOUT, device timeout

---- Unsent message follows ----

Before I start let me admit that I have not tried this. However, I have heard
other people doing it and it works. You can completely dissolve the scotch
tape with acetone. I mean you can dissolve both the adhesive and the backing
in aceotne leaving no residues. I have heard of acetone evaporaotrs where
a little amount of acetone is heated and the vapors condense on the sample.
This acetone dissolves any soluble substance and drips down into the bath to
be reevaporated. The solubility of both adhesive and backing is
higer at near melting point of acetone and hence this process substantially
speeds up the removal. Hope this helps.

Naresh Shah
University of Kentucky
233 Mining & Mineral Bldg.
Lexington, KY 40506-0107
(606)257-4045
naresh-at-funky.mm.uky.edu




From: rms-at-vax.ox.ac.uk
Date: Mon, 29 Nov 1993 14:42:56 +0000
Subject: MICRO 94 2ND CIRCULAR

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

WISSE-at-CYTO.VUB.AC.BE, UHAGOOL-at-rhbnc.ac.uk, SPT1-at-vaxa.york.ac.uk,
SJE-at-PO.CWRU.EDU, SHANTI-at-VISION.EE.UTEXAS.EDU, SFB13-at-phx.cam.ac.uk,
SEEMAN-at-ACFCLUSTER.NYU.EDU, SECR-at-CWI.NL, SCIA-at-CONAN.UIT.NO,
SCHWARZ3-at-URZ.UNIBAS.CH, SCHATT-at-EMAIL.TUWIEN.AC.AT, RYDMARK-at-MEDNT2.SUNET.SE,
RUSS-at-MAT.MTE.NCSU.EDU, RUDOLFO-at-MBL.EDU, ROYSAM-at-ECSE.RPI.EDU,
RMS-at-vax.ox.ac.uk, RIGAUT-at-ARGOS.B3E.JUSSIEU.FR, RICHARD-at-mrcvax.ed.ac.uk,
RETEP-at-ANAT.UCT.AC.ZA, RBRADY-at-NCSA.UIUC.EDU, R.COLEMAN-at-ic.ac.uk,
POSTDO-at-DUTTNCB.TN.TUDELFT.NL, POPOV-at-BIOENG.WASHINGTON.EDU,
PMWOJNAR-at-PLKRCY11.BITNET, PETERS-at-MBCG.UCHC.EDU, PE13-at-phx.cam.ac.uk,
PATERSON-at-UTKVX.UTK.EDU, P.J.KNIGHT-at-bristol.ac.uk, ORMEROD-at-VAXA.ICF.BN.AC.UK,
NICHOLSN-at-i1.ph.gla.ac.uk, MULVEYT-at-spock.vax.aston.ac.uk,
MSHIVAC-at-PEG.PEGASUS.OZ.AU, MONTES-at-VM.CI.UV.ES, MJBACCALA-at-UCDAVIS.BITNET,
MICROSCOPY-at-ANLEMC.MSD.ANL.GOV, MBGEORGE-at-VM.UOGUELPH.CA,
LVJAT-at-siva.bristol.ac.uk, LDP-at-IVEM.MED.UPENN.EDU, LCRUZ-at-ANA.UNIBE.CH
X-VMSmail-To: -at-LIST
Message-ID: {0097644E.7E1F81EA.31927-at-vax.ox.ac.uk}

MICRO 94

International Microscopy and Image Analysis
Conference and Exhibition

12 - 15 September 1994
Earls Court Park Inn, Lillie Road, London

Organized by the Royal Microscopical Society
in association with Microscopy and Analysis

Second Circular

Dates

Conference: Monday 12 September - 2.00 pm - 4.15 pm
Tuesday 13 September - 10.00 am - 4.15 pm
Wednesday 14 September - 10.00 am - 4.15 pm
Thursday 15 September - 10.00 am - 4.15 pm

Exhibition: Monday 12 September - 2.00 pm - 7.00 pm
Tuesday 13 September - 9.30 am - 6.00 pm
Wednesday 14 September - 9.30 am - 6.00 pm
Thursday 15 September - 9.30 am - 4.30 pm

On the Monday evening there will be a wine reception at 5.30pm, followed by
the AGM and Presidental Lecture at 7.00pm.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ


CONFERENCE

Scientific Programme
The Conference will be run in seven half-day sessions. The Electron
Microscopy and Analysis Group of the Institute of Physics (EMAG) and The
Physiological Society are each sponsoring separate lectures within the
Conference.

The Programme will consist of tutorial lectures and posters and will feature
the following topics:-

Monday 12 September (pm) - Materials I
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
EM in the assessment of semiconductor epitaxial growth
Reactions to the surface of implanted bioceramics
X-Ray microanalysis in biomaterials
Optically active nanostructured materials
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Tuesday 13 September (am) - Materials II (including EMAG)
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Auger electron spectroscopy
Imaging time of flight SIMS
Formation of strained layer superlattices by phase separation
Electron microscopy of weakly ordered III-V semiconductor materials
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Tuesday 13 September (pm) Scanning Probe Microscopy
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Applications of atomic force microscopy to thin film research and technology
Scanning probe microscopy: near field imaging of surfaces using electrons,
forces and photons
SPM of living biological systems
Environmental SEM
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Wednesday 14 September (am) - Image Processing and Analysis
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Sampling in 3D for quantitative microscopy
Digital image processing techniques
Image analysis of multicoloured biological specimens
In vivo microscopy by video imaging
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Wednesday 14 September (pm) - 3D Microscopy
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Supercomputing in confocal microscopy
3D atomic-scale microanalysis of materials
Spatial distribution of fibres in composite materials
Confocal polarised-light microscopy
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Thursday 15 September (am) - Flow Cytometry and Proliferation Markers
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Cell cycle control
Proliferation-related proteins
Proliferation in human tumours
Apoptosis
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Thursday 15 September (pm) - Living Cell Cytochemistry (including The
Physiological Society)
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Living cell cytochemistry: Ratio Imaging
Living cell cytochemistry: Confocal scanning laser microscopy

Thursday 15 September (pm) - Proteases in (patho) physiological processes

Use of selective protease inhibitors in the study of collagen breakdown
Role of proteases in invasion and metastasis of cancer cells, arthritis and
rheumatism and infections
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

Experts in the fields listed above have been invited to give lectures. Each
speaker will provide a review of the particular topic in question and ample time
for discussion will be provided.

In addition to the above, a special session will be held on the afternoon of
Monday 12 September on how to use the light microscope.

Technical Lectures will be organized by Exhibitors to act as a bridge between
the specialized review lectures and the equipment being exhibited.
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

INVITED SPEAKERS

It is hoped that the following people will be presenting papers at the MICRO 94
Conference:-

Dr Paul D. Brown (University of Cambridge)
Professor Peter Marquis (University of Birmingham)
Dr Henk Koerten (University of Leiden, The Netherlands)
Dr Peter Dobson (University of Oxford)
Dr R. K. Wild (University of Bristol)
Dr Paul Denison (University of Sheffield)
Dr Andrew Norman (Imperial College, London)
Dr Caroline Baxter (University of Cambridge)
Dr Alan Pidduck (RSRE, Malvern)
Dr M. Miles (HH Wills Physics Laboratory, Bristol)
Dr H Ho”rber(European Molecular Biology Laboratory, Heidelberg, Germany)
Mr Chris Gilpin (Manchester Biological EM Centre)
Dr Vyvyan Howard (Royal Liverpool Children's Hospital)
Dr Dominique Jeulin (Ecole Nationale Superiere des Mines, Fontainebleau,
France)
Dr Hans Tanke (University of Leiden, The Netherlands)
Dr Andreas Kriete (Der Justus Liebig Universitat, Germany)
Dr Alfred Cerezo (University of Oxford)
Dr A. R. Clarke (University of Leeds)
Dr Alan Entwistle (Ludwig Institute for Cancer Research, London)
Dr A Bagg (TNO Rijswijk, The Netherlands)
Dr Michael Ormerod (Sutton, Surrey)
Dr Peter van Mier (Washington University School of Medicine, USA)
Professor P. A. McNaughton (King's College London)
Dr R. Jacob (King's College London)
Dr Vincent Everts (University of Amsterdam, The Netherlands)
Dr Ron Van Noorden (University of Amsterdam, The Netherlands)

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

CONTRIBUTED PAPERS
In addition to invited papers, contributions are invited on all aspects of
microscopy and related techniques.

All contributed papers will appear in the poster sessions of the Conference.
Time will be allowed in the Programme for the viewing of posters, and posters
will be on display for the maximum time possible. At certain times authors will
be in attendance by their posters to discuss their work.

Camera-ready sheets and instructions for the submission of short abstracts
can be obtained from the Royal Microscopical Society office. The deadline for
submission is 4 May 1994. These abstracts will appear in the Conference
Programme, which will be published in a special MICRO 94 issue of the
Proceedings of the Royal Microscopical Society.

Authors will be notified regarding acceptance of their papers by the end of June
1994.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

EXHIBITION
An Exhibition of the latest microscopes and ancillary instrumentation and
equipment will be held at the Earls Court Park Inn, adjacent to the Lecture
Theatre. Admission to the Exhibition is free, by conference badge, or by
exhibition only badge which will be obtainable at the registration desk.

By 1 November 1993, the following firms had reserved exhibition space:-

Agar Scientific Ltd
Alrad Instruments Ltd
Bemax (UK) Ltd
Bio-Rad Laboratories Ltd
British BioCell International
Burleigh Instruments (UK) Ltd
Cambridge Scanning Co Ltd
Confocal Technologies Ltd
Cryophysics Ltd
Data Cell Ltd
Drukker International
Edwards High Vacuum International
Emitech Ltd
Finlay Microvision Co Ltd
Fisons Instruments
Foster Findlay Associates Ltd
Hamamatsu Photonics UK Ltd
Hitachi Scientific Instruments
ISS
Imaging Associates Ltd
J K Instruments Ltd
JEOL UK Ltd
K E Developments Ltd
Lasertec Corporation
Leica Cambridge Limited
Leica UK Limited
Microfield Scientific Ltd
Microscopy and Analysis
Newport Ltd
Nikon UK Limited
Olympus Optical Co (UK) Ltd
Oxford Instruments Microanalysis Group
Oxford Instruments
Philips Electron Optics
Photonic Science
Polaroid (UK) Ltd
Princeton Gamma-Tech (UK) Ltd
Pyser (Holdings) plc
Synoptics Ltd
Taab Laboratories Equipment Ltd
Tracor Europa
Carl Zeiss (Oberkochen) Ltd
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

RECEPTION AND ASSOCIATED EVENTS
On Monday 12 September there will be a wine reception in the Exhibition
between 5.30 pm and 7.00 pm. The Annual General Meeting of the Royal
Microscopical Society, and the Presidential Address will also take place
during the evening.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
ACCOMMODATION
Academic Accommodation
A limited amount of academic accommodation has been booked for the use of
delegates at Imperial College. Rooms have been booked there on a bed and
breakfast basis at a cost of œ2.00 pounds6 per night. This academic/student
accommodation will be filled on a 'first-come first-served' basis. From the
nearby Underground Station at South Kensington, Earls Court is two stops
along the District or Piccadilly Line.

Hotel Accommodation
There are some rooms available in the Earls Court Park Inn at the special
MICRO 94 rate of œ65.00pounds per night. If you would like to reserve
accommodation at these special rates, please contact the Earls Court Park Inn
directly, quoting that you are a MICRO 94 visitor. To be sure of booking rooms
at these rates, it is advisable that you book well in advance.
Telephone: 071 385 1255
Telex: 917728
Fax: 071 381 4450.

Other Hotel Accommodation
Delegates who wish to make their own accommodation arrangements may
wish to use the services of Expotel Executive Travel - Europe's leading hotel
booking agent, who have been appointed the official hotel agency for MICRO
94. The hotel of your choice or a similar alternative can be booked through
Expotel often at discounted rates. By making one telephone call to Expotel on
071 735 0060 stating the event code 'MICRO 94', your reservation will be
confirmed verbally followed by confirmation in writing. This free booking
service is available to anyone attending MICRO 94.
Telephone: 071 735 0060
Telex: 8811951 EXPOTL G
Fax: 071 735 2839.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

REGISTRATION AND PAYMENT

Registration
Registration will take place at the Earls Court Park Inn from 1.00 pm on
Monday 12 September 1994, and from 9.00 am on subsequent mornings.

Payment
Payment may be made by sterling cheque payable to the Royal Microscopical
Society (please add œ12.00pounds to cover exchange and bank charges if
the cheque is not a UK Bank Cheque) or by credit card (Visa/Barclaycard or
Access/Eurocard/Mastercard).

Cancellation and Refunds
Cancellations received before 12 July 1994 will be subject to a full refund. No
refunds will be made if cancellation is made after this date.
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

HOW TO GET THERE

The Conference, Exhibition, Posters, and Refreshments will all take place at
the Earls Court Park Inn, Lillie Road, London SW6.

***************************************************************************


MICRO 94 is organized by the Royal Microscopical Society, 37/38 St Clements,
Oxford OX4 1AJ, UK, in association with Microscopy and Analysis.
Telephone: 0865 248768, Fax: 0865 791237, Email: RMS-at-UK.AC.OX.VAX.




From: Greg Erdos ICBR EM Core Lab University of Florida
Date: Tue, 30 Nov 1993 16:45:57 -0500 (EST)
Subject: NEW LAB

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

{GWERDOS-at-gnv.ifas.ufl.edu}

I am now in the process of designing a new EM suite dedicated to biological
EM. I would like input from those of you who work in what they feel is an
ideal or close to ideal physical arrangement.
I would also like to hear from those who have envisioned an ideal
EM suite but have never seen it built. The restrictions are that a space of
2200 sq ft should house 2 TEMs and 2 SEMs as well as a prep lab, darkrooms
ancillary equipment and 2 offices. The overall design is of more interest
than actual use of the stated space.
So I would like to receive copies of plans you might have or rough
* Director, ICBR EMCL * Any and all contributions would be
greatly appreciated.
*****************************
* Greg Erdos *
* ICBR EM Lab *
* 218 Carr Hall *
* University of Florida *
* Gainesville, FL 32611 *
* gwerdos-at-gnv.ifas.ufl.edu *
* 904-392-1295 *
*****************************




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 30 Nov 1993 20:28:18 -0600 (CST)
Subject: Conference Announcement:Image Processing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 30 Nov 1993 20:41:55 -0600 (CST)
Subject: Abstracts of JOM Articles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


All Subscribers:

Here's yet another addition to the EMail server.
Periodic updates of upcoming Journal articles. Both the
Journal of Microscopy and the Microscopy Society of America
(MSA) Bulletin have agreed in principle to forward this (abstract)
information to me and I will post it to the subscribers.
In many cases the information distributed here may preceed
the actual publication of the full article by a few weeks.

Hope you find it informative! I certainly did.

Nestor Z.
ANL EM Center
=============================================================


JOURNAL OF MICROSCOPY - ABSTRACTS OCTOBER 1993 - DECEMBER 1993

Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 3--12.
Quantitative acoustic microscopy of individual living
human cells
by G. A. D. BRIGGS,* J. WANG* and R. GUNDLE,
*Department of Materials, University of Oxford, Oxford
OX1 3PH, U.K. Nuffield Department of Orthopaedic
Surgery, University of Oxford, Oxford OX3 7LD, U.K.

Summary
The elastic properties of cells can be measured with
microscopic resolution by acoustic microscopy. By
measuring the waveform of very short pulses, the
thickness, and the acoustic velocity, impedance and
attenuation can be determined from the two separate
signals reflected from the top and the bottom of the
cell.




Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 13--21.
Random models for morphological analysis of powders
by D. JEULIN, Centre de Geostatistique, E.N.S.M.P., 35
rue St-Honore, 77305
Fontainebleau, France

Summary
Models to estimate powder size distributions or
fractions of components in a mixture of powders were
developed and tested for various specimens. The models
derived from the Boolean model and from the dead leaves
models can be implemented on rough secondary electron
scanning electron microscope images, obtained after
minimal sample preparation. With the dead leaves
tessellation, the size distributions of spherical
particles or short fibres can be estimated either from
the size distribution of intact grains, or from the
area fraction measurement after binary erosions. With
the dead leaves random function, there is no need for
image segmentation. It provides data on the size and
shape of a population of particles from an estimation
of the distribution of grey-level images after erosions
by convex structuring elements of increasing size. A
version of these models for long fibres is developed
for estimating their diameter distribution. Allowing
for superposition of particles, the proposed methods
enable an unbiased estimation of the size distribution
and characterization of the shape of complex particles.
The approach is illustrated by applications to
spherical particles obtained by simulations and from
scanning electron microscope micrographs.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 23--29.
Spatial distribution of curve length: concept and
estimation
by N. ROBERTS and L. M. CRUZ-ORIVE*, Magnetic Resonance
Research Centre, PO Box 147, University of Liverpool,
Liverpool L69 3BX, U.K. *Stereology Unit, Department of
Anatomy, University of Berne, Postfach 139, CH-3000
Berne 9, Switzerland

Summary
The length of a curvilinear feature, such as a dendrite
tree of a neuron, can, in principle, be estimated by
the recent, non-invasive method of total vertical
projections (TVPs). Curve length is a measure of size,
but it reports nothing about curve shape. The shape of
a tree-like structure can be described to some extent
by the distribution of branch length in properly
defined regions of three-dimensional (3-D) space. A
definition of curve length distribution in three
dimensions is proposed and implemented here on a human
neuron. The relevant 3-D regions overlap after
projection, and therefore the TVPs method cannot be
used directly to estimate the corresponding feature
lengths. However, using the ANALYZE software system
running on a SUNSPARC workstation, dendrite subsets
sitting in predefined regions of space were rendered in
different colours and measured separately by the TVPs
method using a cycloid test system. In combination with
non-invasive image acquisition and processing
techniques, the length distribution concept is likely
to be useful in the metrical analysis of either
microscopic or macroscopic arborizations in a wide
variety of contexts, including living cells and
organisms.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 31--39.
Image sharpness and contrast transfer in coherent
confocal microscopy
by R.OLDENBOURG,* H. TERADA,~~ R. TIBERIO## and S.
INOUE*, *Marine Biological Laboratory, Woods Hole, MA,
U.S.A. Martin Fisher School of Physics, Brandeis
University, Waltham, MA, U.S.A. ~~Hamamatsu Photonics
Kabushiki Kaisha, Hamamatsu City, Japan. ##National
Nanofabrication Facility, Cornell University, Ithaca,
NY, U.S.A.

Summary
Confocal microscopes provide clear, thin optical
sections with little disturbance from regions of the
specimen that are not in focus. In addition, they
appear to provide somewhat greater lateral and axial
image resolution than with non-confocal microscope
optics. To address the question of resolution and
contrast transfer of light microscopes, a new test
slide that enables the direct measurement of the
contrast transfer characteristics (CTC) of microscope
optics at the highest numerical aperature has been
developed. With this new test slide, the performance of
a confocal scanning laser microscope operating in the
confocal reflection mode and the non-confocal
transmission mode was examined. The CTC curves show
that the confocal instrument maintains exceptionally
high contrast (up to twice that with non-confocal
optics) as the dimension of the object approaches the
diffraction limit of resolution; at these dimensions,
image detail is lost with non-confocal microscopes
owing to a progressive loss of image contrast.
Furthermore, we have calculated theoretical CTC curves
by modelling the confocal and non-confocal imaging
modes using discrete Fourier analysis. The close
agreement between the theoretical and experimental CTC
curves supports the earlier prediction that the
coherent confocal and the incoherent non-confocal
imaging mode have the same limit of resolution (defined
here as the inverse of the spatial frequency at which
the contrast transfer converges to zero). The
apparently greater image resolution of the coherent
confocal optics is a consequence of the improved
contrast transfer at spacings which are close to the
resolution limit.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 41--48.
A light-emitting diode light standard
for photo- and videomicroscopy
by J. M. BEACH* and B. R. DULING, Departments of
*Biomedical Engineering and Physiology, University of
Virginia, Charlottesville, VA, U.S.A.
Summary
A light calibration system consisting of a compact
light-emitting diode (LED) source with feedback control
of intensity is described. The source is positioned in
the focal plane of the microscope objective and
produces flat-field illumination of up to 31microW. The
source can be easily used to determine the performance
of microscope optics and camera response. It can also
be used as a standard light source for calibration of
experimental systems. Selectable light intensities are
produced by controlling the LED input power via a
feedback circuit consisting of a photodiode that
detects output light intensity. Spectral coverage
extends between 550 and 670nm using green, yellow and
red LEDs mounted side by side, which are selected
individually. The LED chips are encapsulated in plastic
diffusers which homogenize the light, and a flat field
of illumination is obtained through a thin
1-mm-diameter aperture positioned directly over each
chip. Provision is made for insertion of Ronchi rulings
over the aperture to enable measurements of contrast
modulation in a uniform field. The light may be
pulse-modulated to assess camera response times and the
device can be synchronized with video frames. Narrow
bandpass interference filters can be placed between the
objective lens and the LED source to produce
monochromatic light without affecting the spacing of
controlled light intensities since emission spectra do
not shift appreciably over the range of LED powers
chosen in this design. Results of tests using
controlled light intensity and uniform illumination are
presented.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 49--54.
Video camera calibration for optical densitometry
by R. A. Baldock and I. Poole, MRC Human Genetics Unit,
Crewe Road, Edinburgh EH4 2XU, U.K.

Summary
An efficient technique for calibrating video cameras to
record optical density (OD) from microscopic images is
described. The method corrects for variation over the
field of the brightfield and darkfield intensities,
does not assume a linear response of the camera to the
incident intensity and requires a single calibration
filter.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 55--61.
Vitrification depth can be increased more than 10-fold
by high-pressure freezing
by N. SARTORI, K. RICHTER* and J. DUBOCHET, Laboratoire
d'Analyse Ultrastructurale, Batiment de Biologie,
Universite de Lausanne, CH-1015 Lausanne, Switzerland

Summary
Biological specimens prepared for cryoelectron
microscopy seem to suffer less damage when they are
frozen under 2kbar pressure rather than under normal
conditions. The volume that can be well preserved is
larger. This fact has been illustrated in a number of
publications on a number of different samples. However,
there is a lack of quantitative data concerning the
depth of this good specimen preservation. Catalase
crystals in various sugar solutions have been used as
test objects and vitrification, as determined by
electron diffraction, has been used as the criterion
for good freezing. Keeping all other conditions equal,
the depth of vitrification is approximately 10 times
larger with freezing at high, rather than normal,
pressure. The high-pressure vitrification depth in a
15--20% sugar solution averages 200micrometre. Fully
vitrified specimens up to 700micrometre in thickness
are obtained. When crystalline water is observed it is
frequently in the form of high-density ice II, III or
IX. These results are probably also relevant for
typical biological specimens. The advantage of
high-pressure freezing must be balanced by the possible
consequences of a considerably increased cooling time
and by the damage that may be induced by the pressure.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 63--69.
A portable cryo-storage system for low-temperature
scanning electron microscopy, suitable
for international transport of cryo-specimens
by C. E. Jeffree* and P. R. van Gardingen, *University
of Edinburgh, Science Faculty Electron Microscope
Facility, Daniel Rutherford Building, King's Buildings,
Mayfield Road,
Edinburgh EH9 3JH, U.K.

Summary
A cryo-specimen storage system for low-temperature
scanning electron microscopy (LTSEM) specimens is
described, which: liberates multi-specimen experiments
from sampling restrictions imposed by the rate at which
LTSEM specimens can be examined in the SEM; provides
security against experiment loss resulting from
breakdown of the SEM or cryo-system; enables collection
of specimens in the field or in laboratories remote
from the SEM laboratory; and facilitates international
air transport of LTSEM specimens. The com-ponents of
the system, which has a capacity of 98 stub-mounted
specimens, are readily made in a laboratory workshop.
The details of the design may be altered to suit
particular specimen types or experimental approaches.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 71--79.
Empirically determined freezing time for quick-freezing
with a liquid-nitrogen-cooled copper block
by P. H. W. W. BAATSEN, Departement de Physiologie
generale, FYMU, Universite Catholique de Louvain, Ave.
Hippocrate 55, 1200-Brussels, Belgium

Summary
A method is presented to determine freezing time
empirically. The method is based on determining the
amount of stretch of a skinned muscle fibre while it is
being frozen. Freezing time, as determined with this
method, lies in between 0 and 1ms.


Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 81--88.
An application of scanned focused ion beam milling to
studies on the internal morphology of small arthropods
by R. J. YOUNG,* T. DINGLE,* K. ROBINSON and P. J. A.
PUGH, *FEI Europe Ltd, Brookfield Business Centre,
Cottenham, Cambridge CB4 4PS, U.K. British Antarctic
Survey, Natural Environment Research Council, High
Cross, Madingley Road, Cambridge CB3 0ET,
U.K.

Summary
For the first time a scanned focused ion beam of
approximately 50nm diameter has been used to prepare
biological material. Small defined areas of the surface
were removed by ion etching to allow examination of the
underlying structures with a scanning electron
microscope. Different milling procedures were carried
out on two anatomical features in mites of the genus
Halarachne (Halarachnidae: Mesostigmata). In the first,
square holes were milled into the surface of the
peritrematal plate to reveal the structure of the
underlying respiratory peritrematal groove. In the
second, transverse cuts were made across the shafts of
the sensory sensilli which make up the sensory Haller's
organ on tarsus I. This latter procedure revealed
detail both within the core and walls of sensilli.
Details of specimen preparation and milling procedures,
as well as suitability and interpretation of results,
are presented.



Journal of Microscopy, Vol. 172, Pt. 1,
October 1993, pp. 89--92.
Variable-depth electropolishing of TEM samples
by S. W. LEONARD, Department of Physics, Queen's
University, Kingston, Ontario, K7L 3N6, Canada

Summary
A variable-depth electropolishing technique has been
developed for transmission electron microscopy samples
using copper as sample material. This was required for
an experiment concerning the measurement of the
variation of dislocation density with depth for
ion-implanted materials. The polishing technique was
achieved by a two-step process, involving the
measurement and use of the polishing rate to polish to
a specific depth and the application of a transparent
cover to one side of the sample for back-thinning. With
this technique, any sample depth can be made accessible
for observation with a transmission electron microscope
and the method should be applicable to many different
materials and electropolishers.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 97--107.
Computer simulation of a mirror STEM
by A. V. CREWE, Department of Physics and the Enrico
Fermi Institute, The University of Chicago, 5640 S.
Ellis, Chicago, IL 60464, U.S.A.

Summary
The results of a computer simulation indicate that it
is possible to design and build an STEM that is free
from spherical aberration and should therefore have a
very high resolution. The computer program was written
in APL. The calculations include the effects of
apertures and, consequently, mimic a realistic
situation.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 109--119.
The effect of soft X-radiation on myofibrils
by P. M. BENNETT,* G. F. FOSTER, C. J. BUCKLEY and R.
E. BURGE, *MRC Muscle and Cell Motility Unit, The
Randall Institute, King's College London, 26/29 Drury
Lane, London WC2B 5RL, U.K. Wheatstone Physics
Laboratory, King's College London, The Strand, London
WC2R 2LS, U.K.

Summary
Myofibrils, the contractile organelles from striated
muscles, have been examined in the X-ray microscope to
determine the effect of radiation on their function and
structure. Using X-rays of energy 350--385eV in the
water window we find that after an exposure to 7.5 x
(10 to the fifth power) photons per square micrometre
(calculated to give an absorbed dose of 20 000 Gy) the
myofibrils will no longer contract. The use of the free
radical scavenging agent, DMSO, gives some protection
to the fibrils. It has also been found that after this
much irradiation the fibrils lose up to 20% of their
mass. Further substantial mass loss occurs on
subsequent irradiation. After 25 times the
loss-of-function exposure only 30% of the mass remains.
Analysis of a series of images of the same myofibril
covering this range of exposures shows that the mass is
preferentially lost in some areas of the structure and
consequently significant structural changes occur.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 121--129.
Scanning luminescence X-ray microscopy: imaging
fluorescence dyes at suboptical resolution
by C. JACOBSEN, S. LINDAAS, S. WILLIAMS and X. ZHANG,
Department of Physics, State University of New York at
Stony Brook, Stony Brook, NY 11794, U.S.A.

Summary
Scanning luminescence X-ray microscopy is based on the
use of the very small focused probe of a scanning X-ray
microscope to stimulate visible light emission from
phosphors and dyes. Using an undulator X-ray source and
a Fresnel zone plate to produce a focused X-ray probe,
images of P31 phosphor grains with a resolution of
50--75nm have been obtained, and luminescence from
polystyrene spheres loaded with 50--100micromol/g of
fluorescent dye has been imaged. The resolution was not
limited by the focused X-ray probe (the microscope has
imaged features at 36-nm spacing in transmission mode)
but by dark noise and the low net efficiency of the
luminescence detection system used for this
investigation. This technique may make it possible to
image dye-tagged sites of biochemical activity at the
resolution of the X-ray microscope in wet, unsectioned,
and unfixed cells, especially with soft X-ray optimized
dyes. Because the image is formed from the detection of
signal against a dark background, calculations suggest
that the radiation dose for luminescence imaging of
dye-tagged features should be 2--22 times lower than
it is in transmission X-ray microscopy. A possible
extension of the technique for three-dimensional
imaging at the transverse resolution of the X-ray
microscope is described, where visible light collection
optics might be used to obtain submicrometre axial
resolution.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 131--136.
High-spatial-resolution maps of sulphur from human hair
sections: an EELS study
by P. HALLEGOT and P. CORCUFF, Laboratoires de
Recherche Avancee, L'Oreal, Departement de Biophysique,
1 avenue Eugene-Schueller, 93 600 Aulnay sous Bois,
France

Summary
High-resolution sulphur maps have been acquired from
human hair using a Zeiss CEM 902A transmission electron
microscope equipped with an energy filter. Analysis by
electron energy-loss spectroscopy (EELS) was performed
on ultrathin sections of hair shafts embedded in three
different types of resin: Nanoplast (water-soluble),
Spurr (epoxy) and Lowicryl (low-temperature resin).
Good-quality energy-loss images have been obtained with
the three resins, although it was found that Nanoplast
gave the best image contrast. For the first time, the
results obtained for the detection of sulphur by silver
staining of hair sections, which until now has been the
only way to map sulphur at the electron microscopic
level, have been confirmed. The results are compared
with local sulphur concentrations from bulk analysis.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 137--151.
Improvements in the technique of vascular
perfusion-fixation employing a fluorocarbon-containing
perfusate and a peristaltic pump controlled by pressure
feedback
by J. Rostgaard, K. Qvortrup and S. S. Poulsen,
Institute of Medical Anatomy Department B, The Panum
Institute, University of Copenhagen, 3 Blegdamsvej,
2200 Copenhagen N, Denmark

Summary
A new, improved technique for whole-body
perfusion-fixation of rats and other small animals is
described. The driving force is a peristaltic pump
which is feedback regulated by a pressure transducer
that monitors the blood/perfusion pressure in the left
ventricle of the heart. The primary perfusate-fixative
is composed of a blood substitute---13.3% oxygenated
fluorocarbon FC-75---in 0.05 M cacodylate buffer (pH
7.4) with 2% glutaraldehyde. The secondary
perfusate-fixative is composed of 2% glutaraldehyde in
0.05 M cacodylate buffer (pH 7.4) with 20 mM CaCl2 A
double-barrelled, self-holding cannula is used to
cannulate the heart; the outer and inner barrels of the
cannula are connected to the peristaltic pump and to
the pressure transducer, respectively. The tissue
oxygen tension in the rat is monitored by a
sub-cutaneous oxygen electrode. Measurements showed
that tissue hypoxia/anoxia did not develop before or
during the perfusion-fixation. Thus, the technique
permits study of specimens which do not exhibit
fixation gradients and do not contain cells fixed in a
state of asphyxia. This is substantiated by electron
micrographs of cells from different organs, revealing
new fine structural elements. By adding oxygenated
fluorocarbon to glutaraldehyde perfusate-fixatives,
enough oxygen is made accessible for cellular
respiration as well as for the oxygen-consuming
chemical reactions of glutaraldehyde with the tissue.
Data on anaesthesia, operative manoeuvres, mechanical
components of the system, preparation of fixatives and
flow of the perfusate-fixatives are furnished and
discussed.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 153--156.
A device for picking up ultrathin serial sections
by E. P. Meyer and V. J. Domanico, Department of
Zoology, University of Zurich, Winterthurerstrasse 190,
CH-8057 Zurich, Switzerland

Summary
Reconstruction from serial sections is often required
to understand the architecture of biological tissue.
The sampling of serial sections calls for an enormous
amount of patience and skill. Sections are cut with a
diamond knife, sampled in a trough and must then be
collected on grids for examination with the electron
microscope (EM). This last step, in which the section
ribbons have to be aligned with the EM grid, is
especially difficult as both hands have to work in
perfect co-ordination. A simple manually operated or
motorized mechanical device has been designed which
facilitates the collection of EM section ribbons.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 157--162.
An unbiased and efficient procedure for 3-D
connectivity measurement as applied to porous
media
by H. Q. ZHAO and I. F. MACDONALD, Porous Media
Research Institute, Department of Chemical Engineering,
University of Waterloo, Ontario, Canada N2L 3G1

Summary
A new stereological technique to measure the mean genus
(connectivity) per unit volume of a porous medium is
described and applied to a real sandstone sample. The
technique is based on the `net volume tangent counts'
performed on disector samples, i.e. pairs of
consecutive sections of an interconnected structure. It
consists of a set of simple counting rules applied to
the features on the sections of the structure and can
be easily implemented manually. The applicability and
efficiency of the procedure is evaluated by applying it
to a Berea sandstone sample which has been studied
previously using a network analysis approach by
interactive three-dimensional computer reconstruction.
It is shown that the procedure yields results in good
agreement with the network analysis result, but has the
advantages that it is much easier to implement, is more
flexible in how the data are collected, is more
efficient, and is known to provide an unbiased estimate
of the mean genus per unit volume of the whole
structure.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 163--176.
Remapping disparate images for coincidence
by W. Galbraith* and D. L. Farkas, *Department of
Laboratory Medicine, Allegheny Singer Research
Institute, 320 East North Ave., Pittsburgh, PA 15212,
U.S.A. Center for Light Microscope Imaging and
Biotechnology, Carnegie-Mellon University, 4400 Fifth
Avenue, Pittsburgh, PA 15213, U.S.A.

Summary
With the development of complex multimode computerized
microscope systems, it is possible and necessary to
obtain images of the same area of the microscopical
preparation by several methods of microscopy, such as
differential interference contrast, reflection
interference microscopy, several wavelengths of
fluorescence microscopy, laser scanning and confocal
modes. Thus, varied information may be obtained about a
single field, in the form of a set of images, taken at
different ports of the microscope, using different
digitizing cameras, each appropriate to certain tasks.
For comparative purposes, the images should be
superimposable, pixel by pixel, but in general they are
not --- they differ in image shape and size,
magnification, distortion, centration and orientation.
This paper shows how the problem may be approached,
using an extension of the remapping procedures
described in a previous paper, in which images of a
separate grid reference slide are used to detect,
quantify and correct the image errors. Affine
remapping, without the use of grid images, is also
described.



Journal of Microscopy, Vol. 172, Pt. 2, November 1993,
pp. 177--180.
Simplified nerve cell counting in the rat brainstem
with the physical disector using a drawing-microscope
by O. GUNTINAS-LICHIUS,* J. MOCKENHAUPT,* E.’STENNERT
and W. F. NEISS*, *Department of Anatomy and Department
of Oto-Rhino-Laryngology, University of Cologne,
Lindenthal, D-50924 Koln, Germany

Summary
A simple modification of the physical disector is
presented, which is used to count the number of neurons
in the hypoglossal nucleus of the rat in a series of
paraffin sections. One disector consists of two
adjacent sections (6micrometre thick) that have been
Nissl-stained with cresyl fast violet. In the first
step of the procedure each of the two sections is
investigated separately with a drawing-microscope. The
boundary of the hypoglossal nucleus and the position of
neurons devoid of, or containing a part of, the cell
nucleus in the plane of the section are marked on
transparent paper. In the second step, these two
drawings are placed one upon another, aligned and the
number of cell profiles that show a cell nucleus in
one but not in both drawings counted. This modification
of the disector method for cell counting needs no
specialized equipment, simply a light microscope with
drawing apparatus, and can be combined with
histochemical studies of other sections from the same
tissue block.




Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 189--194.
A compact Schwarzschild soft X-ray microscope with a
laser-produced plasma source
by Y. HORIKAWA, K. NAGAI, S. MOCHIMARU and Y. IKETAKI,
T. Morokuma Research Laboratory, Olympus Optical Co.,
Ltd, 2--3 Kuboyama-cho, Hachioji-shi, Tokyo 192, Japan

Summary
A compact Schwarzschild soft X-ray microscope using a
laser-produced plasma soft X-ray source has been
developed. The laser-produced plasma source, which is
small but of high brilliance, has made it possible to
use the soft X-ray microscope in a small laboratory.
The microscope is composed of a Schwarzschild objective
and a grazing incidence mirror condenser. Image
contrast for biological specimens in soft X-ray regions
is investigated briefly. It is possible to observe the
fine structures of a thin specimen at a wavelength of
15nm; at this wavelength high-contrast images of
biological specimens have been obtained with a single
laser shot of pulse width of 8ns at a resolution of
0.3micrometre. The resolution of the system is limited
by the detector.



Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 195--203.
Liquid substitution: a versatile procedure for SEM
specimen preparation of biological materials without
drying or coating
by H. J. Ensikat and W. Barthlott, Botanisches Institut
der Universitat, Meckenheimer Allee 170, 53115 Bonn,
Germany

Summary
Certain liquids with a very low vapour pressure, such
as glycerol or triethylene glycol, can be used to
infiltrate biological specimens so that they may be
observed in the scanning electron microscope (SEM)
without drying. The conductive properties of the fluids
allow specimens to be examined either uncoated or with
very thin coatings. The advantages of liquid
substitution include the retention of lipids, waxes,
loose particles, and surface contaminants. Since the
procedure does not require expensive equipment, it
offers an alternative to critical point drying or
cryo-preparation. For certain types of specimens,
liquid substitution may represent the best preparation
procedure. In addition, the fluids themselves may be
imaged directly in the SEM, or indirectly by
cathodoluminescence following labelling with
fluorochromes.



Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 205--214.
Conformational characterization of nucleosomes by
principal component analysis of their electron
micrographs
by M. M. Z. ZABAL*, G. J. CZARNOTA*, D. P. BAZETT-JONES
and F. P.
OTTENSMEYER*, *Department of Medical Biophysics, The
University of Toronto and Ontario Cancer Institute, 500
Sherbourne Street, Toronto, Ontario, Canada M4X 1K9

Summary
Optimized fixation conditions were determined for
protein--protein and protein--DNA crosslinking within
calf-thymus nucleosomes in low monovalent salt
concentrations. Nucleosomes were examined without
heavy-atom staining by darkfield electron microscopy.
The dimensions of these macromolecular complexes and
those of HeLa core particles optimally fixed in
divalent salt were analysed using principal component
analysis. According to this analysis the structure of
the calf-thymus nucleosomes was best presented by a
prolate ellipsoid. Particle images had average major
and minor axis lengths of 14.1 and 10.5nm,
respectively. In contrast, the HeLa nucleosomes were
best modelled by an oblate ellipsoid from the analysis
of their images, which had average major and minor axes
of 13.3 and 11.5nm. The applicability of this
multivariate statistical analysis to the interpretation
of macromolecular images is illustrated and discussed.



Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 215--221.
An investigation of substrate and sample preparation
effects on scanning tunnelling microscopy studies on
xanthan gum
by M. J. WILKINS, M. C. DAVIES, D. E. JACKSON, C. J.
ROBERTS and S. J. B. TENDLER, The Laboratory of
Biophysics and Surface Analysis, Department of
Pharmaceutical Sciences, The University of Nottingham,
University Park, Nottingham NG7 2RD, U.K.

Summary
Scanning tunnelling microscopy is developing as an
important biophysical tool for the molecular imaging of
biological material. In this study, the effect of
sample deposition technique and substrate employed on
the resultant images of a microbial polysaccharide is
investigated. Scanning tunnelling microscopy topographs
of xanthan gum, pipette deposited and spray deposited
onto highly orientated pyrolytic graphite and mica
substrates are presented. The use of pipette deposition
of the aqueous sample solution is shown to result in a
xanthan network. In contrast, images of isolated
xanthan molecules are obtained when spray deposition
with glycerol is employed. The effect of these
deposition techniques on the macroscopic distribution
of sample material across substrates is shown using
confocal fluorescence microscopy.


Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 223--232.
Fractal characterization by frequency analysis. I.
Surfaces
by E. ANGUIANO, M. PANCORBO and M. AGUILAR, Instituto
de Ciencia de Materiales, Sede B (CSIC), Universidad
Auto¨noma de Madrid (C-III), 28049-Madrid, Spain

Summary
A study of the quality and accuracy of the methods
based on frequency analysis for the fractal
characterization of surfaces as measured by scanning
tunnelling microscopy (or profilometry) is made. The
study is based on computer simulation of images of
fractal surfaces. A discussion of the mathematical
algorithms used for computer generation of fractal
surfaces then follows. The main conclusion is that
studies of fractal characterization by frequency
analysis reported in previous papers in the STM field,
as well as conclusions about the performance of the
various methods, are doubtful. New methods for
frequency analysis that in some cases produce more
reliable results are proposed.



Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 233--238.
Fractal characterization by frequency analysis. II. A
new method
by M. AGUILAR, E. ANGUIANO and M. PANCORBO, Instituto
de Ciencia de Materiales, Sede B (CSIC), Universidad
AutÆnoma de Madrid (C-III), 28049-Madrid, Spain

Summary
A new frequency analysis method, fractal analysis by
circular average (FACA), and an image replication
procedure are proposed that together produce accurate
measurements of the fractal dimension of surfaces and
profiles, eliminating Fourier transform artefacts which
arise from the lack of periodic continuity in real
surfaces and profiles.



Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 239--248.
Effects of noise and anisotropy on the determination of
fractal dimensions
by J. C. RUSS, Materials Science and Engineering
Department, North Carolina State University, Raleigh,
NC 27695-7907, U.S.A.

Summary
Measurement of the fractal dimension of surfaces imaged
by the scanning tunnelling microscope, atomic force
microscope or similar instruments can be performed
using several different algorithms. Dimensional
analysis---plots of log (perimeter) vs. log
(area)---are compared to Korcak and slit-island
methods, which also employ a horizontal plane section,
and to Fourier analysis of the two-dimensional array of
elevation values. The effects of surface anisotropy and
instrument noise on each of these measurement
techniques is investigated.



Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 249--256.
Evaluation of the precision of systematic sampling:
nugget effect and covariogram modelling
by J. THIOULOUSE,* J. P. ROYET, H. PLOYE* and F.
HOULLIER~~, *Laboratoire de Biometrie, Genetique et
Biologie des Populations, (U.R.A.CNRS 243), Universite
Claude Bernard-Lyon 1, F-69622 Villeurbanne Cedex,
France. Laboratoire de Physiologie Neurosensorielle,
(U.R.A. CNRS 180), Universite Claude Bernard-Lyon 1,
F-69622 Villeurbanne Cedex, France. ~~Ecole Nationale
du Genie Rural, des Eaux et Forets, Laboratoire de
Recherches en Sciences forestieres, 14, rue Girardet,
F-54042 Nancy Cedex, France

Summary
Systematic sampling designs are widely used in
stereology. When an estimator of the total amount, Q,
of the sampled variable is evaluated by such a
procedure, the coefficient of error can be predicted by
applying Matheron's theory of regionalized variables.
To evaluate the accuracy of the estimate of Q, it is
necessary to study the behaviour of the regionalized
variable and to model its covariogram. Histological
data with a low short-range variability and agronomic
data with a pronounced nugget effect provided the
biological material for extreme case studies. Results
show that the short-range variability, if present,
cannot be detected when only small samples are
available. An underestimation of the coefficient of
error is then to be expected. We propose several models
of the covariogram, which can be used to test for the
presence of a nugget effect. If a nugget effect is
present, these models will provide better estimates of
the coefficient of error. If there is no nugget effect
a simplified method can be used and will provide
reliable estimates of the coefficient of error.



Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 257--261.
Directional analysis of planar fibre networks:
application to cardboard microstructure
by I. MOLCHANOV,* D. STOYAN* and K.’FYODOROV,
*Bergakademie Freiberg, FB Mathematik,
Bernhard-v.-Cotta-Str. 2, D-09596 Freiberg, Germany.
Kiev Technological Institute of the Food Industry,
Vladimirskaya, 68, 252017 Kiev, Ukraine

Summary
This paper discusses the problem of determining
suitable roses of intersections for systems of planar
thick fibres which cross and overlap. A planar Boolean
model with long rectangular (or similar) grains is
suggested as an appropriate mathematical model. It
leads to a statistical estimator for the rose of
intersections of the system of fibre spines. The method
is used to analyse the microstructure of the outer
layer of two samples of cardboard.


Journal of Microscopy, Vol. 172, Pt. 3, December 1993,
pp. 263--266.
Edge detection in petrographic images
by J. STARKEY and A. K. SAMANTARAY, Department of Earth
Sciences, The University of Western Ontario, London,
Ontario, Canada N6A 3B7

Summary
The automatic detection of mineral grain boundaries in
images obtained from a polarized-light microscope
requires special techniques. Observations in both
plane- and cross-polarized light may be necessry and
the section must be rotated relative to the plane of
polarization of the microscope to see all the grain
boundaries. In computer-based microscopy this can be
accomplished by the sequential accumulation of
individual images captured from one microscope field of
view with different polarizer orientations. ƒFor
real-time implementation the sequential images are
segmented individually by applying Canny's algorithm. A
separable Gaussian mask is used for smoothing and a 3 x
3 convolution mask is used to generate 1-pixel-wide
boundaries, which are located at the zero-crossing of
the second-order derivative of the intensity gradient.
The boundaries are extracted and accumulated in a
composite image. The resulting composite image is a
synoptic grain-boundary image of the rock.




From: {ZALUZEC-at-anlemc.msd.anl.gov}:ddn:wpafb
Date: 11-27-93 12:55pm
Subject: TEM:Celli Tape

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9312010147.AA22920-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: TEM:Celli Tape
Orig-Author: {"Nestor J. Zaluzec (708)-252-5075, -4964" {ZALUZEC-at-anlemc.msd.anl.gov} }:ddn:wpafb
-----------------------------------------------------------
Re: TEM specimen preparation of Graphite:

Here's a question for those of you that used to use the
old Celli tape popular in England about 10+ years ago.

In the past I had some of this tape which I used to use
to prepare specimens of graphite for TEM. Here I would
use the old trick of just continually touching the surface
of the graphite with fresh tape to delaminate the graphite
gradually until I made a nice thin & optically transparent
section, which stuck to the tape and was usually perfect for
TEM. The nice thing about this old type of scotch tape was that
the adhesive used would disolve beautifully in Acetone and the
backing for that tape would just float away as it was not
soluablein the acetone. Unfortuantely, the current type
of "SCOTCH BRAND" transparent tape completely dissolves
in Acetone and most other solvents I tried.

I'm back to trying to make more graphite samples as
my old ones have finally bit the dust so to speak.
Since the new scotch tape failed, I tried extraction
replica tape which is available from most EM supply
vendors, however it does not have the adhearance
of either the old Celli tape or the newer scotch tapes.
(BTW the new scotch tape very nicely delaminated the graphite
and gives beautifully thin sections if I could ever
remove them from the adhesive). Does anyone know of a
solvent which works on the newer scotch brand tapes
or have another/similar idea/procedure for layered
structures which are not strongly bound?.

I'd prefer to avoid ion milling the graphite although
I may have to resort to this in the end.

Nestor Z.
ANL EM Center





From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 30 Nov 1993 20:29:42 -0600 (CST)
Subject: MICRO 94 2ND CIRCULAR

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



MICRO 94

International Microscopy and Image Analysis
Conference and Exhibition

12 - 15 September 1994
Earls Court Park Inn, Lillie Road, London

Organized by the Royal Microscopical Society
in association with Microscopy and Analysis

Second Circular

Dates

Conference: Monday 12 September - 2.00 pm - 4.15 pm
Tuesday 13 September - 10.00 am - 4.15 pm
Wednesday 14 September - 10.00 am - 4.15 pm
Thursday 15 September - 10.00 am - 4.15 pm

Exhibition: Monday 12 September - 2.00 pm - 7.00 pm
Tuesday 13 September - 9.30 am - 6.00 pm
Wednesday 14 September - 9.30 am - 6.00 pm
Thursday 15 September - 9.30 am - 4.30 pm

On the Monday evening there will be a wine reception at 5.30pm, followed by
the AGM and Presidental Lecture at 7.00pm.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ


CONFERENCE

Scientific Programme
The Conference will be run in seven half-day sessions. The Electron
Microscopy and Analysis Group of the Institute of Physics (EMAG) and The
Physiological Society are each sponsoring separate lectures within the
Conference.

The Programme will consist of tutorial lectures and posters and will feature
the following topics:-

Monday 12 September (pm) - Materials I
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
EM in the assessment of semiconductor epitaxial growth
Reactions to the surface of implanted bioceramics
X-Ray microanalysis in biomaterials
Optically active nanostructured materials
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Tuesday 13 September (am) - Materials II (including EMAG)
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Auger electron spectroscopy
Imaging time of flight SIMS
Formation of strained layer superlattices by phase separation
Electron microscopy of weakly ordered III-V semiconductor materials
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Tuesday 13 September (pm) Scanning Probe Microscopy
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Applications of atomic force microscopy to thin film research and technology
Scanning probe microscopy: near field imaging of surfaces using electrons,
forces and photons
SPM of living biological systems
Environmental SEM
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Wednesday 14 September (am) - Image Processing and Analysis
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Sampling in 3D for quantitative microscopy
Digital image processing techniques
Image analysis of multicoloured biological specimens
In vivo microscopy by video imaging
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Wednesday 14 September (pm) - 3D Microscopy
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Supercomputing in confocal microscopy
3D atomic-scale microanalysis of materials
Spatial distribution of fibres in composite materials
Confocal polarised-light microscopy
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Thursday 15 September (am) - Flow Cytometry and Proliferation Markers
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Cell cycle control
Proliferation-related proteins
Proliferation in human tumours
Apoptosis
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Thursday 15 September (pm) - Living Cell Cytochemistry (including The
Physiological Society)
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
Living cell cytochemistry: Ratio Imaging
Living cell cytochemistry: Confocal scanning laser microscopy

Thursday 15 September (pm) - Proteases in (patho) physiological processes

Use of selective protease inhibitors in the study of collagen breakdown
Role of proteases in invasion and metastasis of cancer cells, arthritis and
rheumatism and infections
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

Experts in the fields listed above have been invited to give lectures. Each
speaker will provide a review of the particular topic in question and ample time
for discussion will be provided.

In addition to the above, a special session will be held on the afternoon of
Monday 12 September on how to use the light microscope.

Technical Lectures will be organized by Exhibitors to act as a bridge between
the specialized review lectures and the equipment being exhibited.
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

INVITED SPEAKERS

It is hoped that the following people will be presenting papers at the MICRO 94
Conference:-

Dr Paul D. Brown (University of Cambridge)
Professor Peter Marquis (University of Birmingham)
Dr Henk Koerten (University of Leiden, The Netherlands)
Dr Peter Dobson (University of Oxford)
Dr R. K. Wild (University of Bristol)
Dr Paul Denison (University of Sheffield)
Dr Andrew Norman (Imperial College, London)
Dr Caroline Baxter (University of Cambridge)
Dr Alan Pidduck (RSRE, Malvern)
Dr M. Miles (HH Wills Physics Laboratory, Bristol)
Dr H Ho”rber(European Molecular Biology Laboratory, Heidelberg, Germany)
Mr Chris Gilpin (Manchester Biological EM Centre)
Dr Vyvyan Howard (Royal Liverpool Children's Hospital)
Dr Dominique Jeulin (Ecole Nationale Superiere des Mines, Fontainebleau,
France)
Dr Hans Tanke (University of Leiden, The Netherlands)
Dr Andreas Kriete (Der Justus Liebig Universitat, Germany)
Dr Alfred Cerezo (University of Oxford)
Dr A. R. Clarke (University of Leeds)
Dr Alan Entwistle (Ludwig Institute for Cancer Research, London)
Dr A Bagg (TNO Rijswijk, The Netherlands)
Dr Michael Ormerod (Sutton, Surrey)
Dr Peter van Mier (Washington University School of Medicine, USA)
Professor P. A. McNaughton (King's College London)
Dr R. Jacob (King's College London)
Dr Vincent Everts (University of Amsterdam, The Netherlands)
Dr Ron Van Noorden (University of Amsterdam, The Netherlands)

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

CONTRIBUTED PAPERS
In addition to invited papers, contributions are invited on all aspects of
microscopy and related techniques.

All contributed papers will appear in the poster sessions of the Conference.
Time will be allowed in the Programme for the viewing of posters, and posters
will be on display for the maximum time possible. At certain times authors will
be in attendance by their posters to discuss their work.

Camera-ready sheets and instructions for the submission of short abstracts
can be obtained from the Royal Microscopical Society office. The deadline for
submission is 4 May 1994. These abstracts will appear in the Conference
Programme, which will be published in a special MICRO 94 issue of the
Proceedings of the Royal Microscopical Society.

Authors will be notified regarding acceptance of their papers by the end of June
1994.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

EXHIBITION
An Exhibition of the latest microscopes and ancillary instrumentation and
equipment will be held at the Earls Court Park Inn, adjacent to the Lecture
Theatre. Admission to the Exhibition is free, by conference badge, or by
exhibition only badge which will be obtainable at the registration desk.

By 1 November 1993, the following firms had reserved exhibition space:-

Agar Scientific Ltd
Alrad Instruments Ltd
Bemax (UK) Ltd
Bio-Rad Laboratories Ltd
British BioCell International
Burleigh Instruments (UK) Ltd
Cambridge Scanning Co Ltd
Confocal Technologies Ltd
Cryophysics Ltd
Data Cell Ltd
Drukker International
Edwards High Vacuum International
Emitech Ltd
Finlay Microvision Co Ltd
Fisons Instruments
Foster Findlay Associates Ltd
Hamamatsu Photonics UK Ltd
Hitachi Scientific Instruments
ISS
Imaging Associates Ltd
J K Instruments Ltd
JEOL UK Ltd
K E Developments Ltd
Lasertec Corporation
Leica Cambridge Limited
Leica UK Limited
Microfield Scientific Ltd
Microscopy and Analysis
Newport Ltd
Nikon UK Limited
Olympus Optical Co (UK) Ltd
Oxford Instruments Microanalysis Group
Oxford Instruments
Philips Electron Optics
Photonic Science
Polaroid (UK) Ltd
Princeton Gamma-Tech (UK) Ltd
Pyser (Holdings) plc
Synoptics Ltd
Taab Laboratories Equipment Ltd
Tracor Europa
Carl Zeiss (Oberkochen) Ltd
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

RECEPTION AND ASSOCIATED EVENTS
On Monday 12 September there will be a wine reception in the Exhibition
between 5.30 pm and 7.00 pm. The Annual General Meeting of the Royal
Microscopical Society, and the Presidential Address will also take place
during the evening.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ
ACCOMMODATION
Academic Accommodation
A limited amount of academic accommodation has been booked for the use of
delegates at Imperial College. Rooms have been booked there on a bed and
breakfast basis at a cost of œ2.00 pounds6 per night. This academic/student
accommodation will be filled on a 'first-come first-served' basis. From the
nearby Underground Station at South Kensington, Earls Court is two stops
along the District or Piccadilly Line.

Hotel Accommodation
There are some rooms available in the Earls Court Park Inn at the special
MICRO 94 rate of œ65.00pounds per night. If you would like to reserve
accommodation at these special rates, please contact the Earls Court Park Inn
directly, quoting that you are a MICRO 94 visitor. To be sure of booking rooms
at these rates, it is advisable that you book well in advance.
Telephone: 071 385 1255
Telex: 917728
Fax: 071 381 4450.

Other Hotel Accommodation
Delegates who wish to make their own accommodation arrangements may
wish to use the services of Expotel Executive Travel - Europe's leading hotel
booking agent, who have been appointed the official hotel agency for MICRO
94. The hotel of your choice or a similar alternative can be booked through
Expotel often at discounted rates. By making one telephone call to Expotel on
071 735 0060 stating the event code 'MICRO 94', your reservation will be
confirmed verbally followed by confirmation in writing. This free booking
service is available to anyone attending MICRO 94.
Telephone: 071 735 0060
Telex: 8811951 EXPOTL G
Fax: 071 735 2839.

þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

REGISTRATION AND PAYMENT

Registration
Registration will take place at the Earls Court Park Inn from 1.00 pm on
Monday 12 September 1994, and from 9.00 am on subsequent mornings.

Payment
Payment may be made by sterling cheque payable to the Royal Microscopical
Society (please add œ12.00pounds to cover exchange and bank charges if
the cheque is not a UK Bank Cheque) or by credit card (Visa/Barclaycard or
Access/Eurocard/Mastercard).

Cancellation and Refunds
Cancellations received before 12 July 1994 will be subject to a full refund. No
refunds will be made if cancellation is made after this date.
þÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄÄ

HOW TO GET THERE

The Conference, Exhibition, Posters, and Refreshments will all take place at
the Earls Court Park Inn, Lillie Road, London SW6.

***************************************************************************


MICRO 94 is organized by the Royal Microscopical Society, 37/38 St Clements,
Oxford OX4 1AJ, UK, in association with Microscopy and Analysis.
Telephone: 0865 248768, Fax: 0865 791237, Email: RMS-at-UK.AC.OX.VAX.




From: Nestor J. Zaluzec (708)-252-5075, -4964 :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 30 Nov 1993 21:51:44 -0600 (CST)
Subject: Celli (Sellotape) & cleavage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Problem solved thanks to many suggestions.

Summary of the procedure for cleaving graphite
into TEM specimens using 3M brand Scotch (Magic) tape.

Begin using with the Scotch Brand Magic Tape using it to
cleave the graphite flakes several times to get a "clean"
i.e. fresh surfaces. Then begin with a fresh piece of tape.
Continue to delaminate until you see a number (~10+) of optically
transparent "flakes" stuck onto the tape (many more will not
be optically transparent) . During the delamination
/peeling process try not to press the tape pieces together with
alot of force this only serves to make complications later. Use
only sufficient force to cause the graphite flakes to stick & peel off.
Now trim off excess tape where there is no potential graphite
specimens attached. Immerse the remaining tape in Toluene at room temp ,
this starts to seperate/loosen the backing from the glue, after
about 5 minutes slowly add Methanol until you get to a ~50/50 mixture
Let this sit a few more minutes then gently loosen the glue mass
(looks like a wrinkled gelatinous mass and acts like it too) from the backing .
To remove the plastic backing, you may need to use a pair of tweezers
to entice the seperation where the glue has been forced down
into good contact with the plastic backing at the sites of the
graphite flakes. Toss the plastic backing in the trash.

Now begin to add acetone which dissolves the glue.
As the acetone is added use a pair of
tweezers to agitate the glue/graphite mass. This allows the
solvent to get inbetween the graphite flakes and glue mass. Remove
as much of the glue mass as possible before it completely
dissolves in the solvent mixture. This is a real sticky "goo"
at this point and it will only serve to contaminate your
solution further. Be realistic here you will
not get all of the graphite off of the tape. I estimate that
there were about 2 dozen specimens on the tape, as the graphite
released from the glue it fractured into maybe a hundred smaller
flakes which were just floating in the solvent. At some point
it nolonger is worth while trying to release the remaining graphite flakes
from the glue mass, too much agitation only causes folds which
trap the flakes and makes it nearly impossible to release all
the graphite. Waiting to long only causes the glue mass to
dissolve into your solvent and possibly contaminate your specimens.


Decant off excess solvent and add fresh acetone to continue dissolution
of any remaining glue on the flakes. Now just pickup the flakes on
grids by floating them over grids and lifting out of the solvent.
(i.e. play chase the flake with the grid. As I recall the life science
microscopists have a special tool for this but I just went fishing with a pair
of tweezers and a grid under a low power steroe microscope).
If they (the graphite flakes) are optically
transparent (very light gray to clear) you've
got a reasonably good specimen to work with at 100 kV.

Via TEM I see the usual basal plane dislocations, and little
surface contamination there is some but it is obvious and can
be easily avoided. The flakes show the usual bending that
I would expect from the peeling procedure, but are reasonably
uniform in thickness over wide areas (~sq microns). No thickness
measurements yet, I'll report those when the data comes out
of the experiments.

Thanks to all who made suggestions:

Nestor Z.
ANL EM Center




MicroscopyListserver Archive Email Extraction Software Version NJZ07060908