From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 1 Jun 1994 8:55:00 -0500 (CDT)
Subject: Chemical Shifts in X-ray Spectra

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To continue the Thread....

Although X-ray spectra can do this sort of work, isn't it true
that for the most part the electron spectroscopies tend to do
the job easier? I'm thinking here of mainly of ESCA/XPS (i.e.
the photoelectron spectroscopies).

Nestor Zaluzec
ANL EMCenter




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 1 Jun 1994 8:58:51 -0500 (CDT)
Subject: Screens:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Pierre

Once you have collected all the wisdom of the world on Screens
please post a summary here for all to read.

The same should apply to everyone that posts a question and collects
answers. Sometimes as Pierre pointed out not all replies go to the
listserver and you would be "paying your dues" by providing a
concise summary of all reponses for all subscribers to read and
possibly file for future use.

Nestor Zaluzec
ANL EMCenter




From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 1 Jun 1994 12:23:54 U
Subject: Ion Tech Address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subject: Time:12:22 PM
OFFICE MEMO Ion Tech Address Date:6/1/94
I have an address for Ion Tech Inc. at 2330 East Prospect,
Ft. Collins, CO, 80525. Tel: 303-221-1807; Fax: 303-493-1439.
My source gives offices in Germany and Japan, but none in the UK,
so this may not be the organization you are interested in.






From: bnhirsch-at-wicc.weizmann.ac.il (David L. Hirschberg)
Date: Thu, 2 Jun 1994 11:15:16 +0300
Subject: LM Fluorescence: Nuclear stain that can be evaluated with filters for FITC or Rhodamine

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199406020813.LAA04376-at-wisslc1.weizmann.ac.il}
X-Sender: bnhirsch-at-wisslc1.weizmann.ac.il
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Is there fluorescent stain that can be applied as easily as DAPI or
Hoescht stains that have excitations and emissions in the same
wavelength as FITC or Rhodamine?

Thanks, David

=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=-=
David L. Hirschberg bnhirsch-at-wicc.weizmann.ac.il
Department of Neurobiology (972) (0)834-2127 (0)834-2412 work
Weizmann Institute of Science (972) 847-4805 home
Rehovot 76100 Israel (972) 834-4131 fax






From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Thu, 2 Jun 1994 08:30:58 +0000 (GMT)
Subject: EM/XPS : Re: Chemical shifts in x-ray spectra

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {20839.9406020831-at-irix.bris.ac.uk}

} Although X-ray spectra can do this sort of work, isn't it true
} that for the most part the electron spectroscopies tend to do
} the job easier? I'm thinking here of mainly of ESCA/XPS (i.e.
} the photoelectron spectroscopies).
}
} Nestor Zaluzec
} ANL EMCenter
}
It depends on the spatial resolution you require. Here in Bristol we have
a VG Escascope small area/imaging x-ray photoelectron spectrometer. The
smallest area you can obtain XPS spectra from is 50um diameter, so that is
the smallest area from which you can obtain chemical shift information (as
long as you are prepared to wait a considerable time for the spectrum to
appear out of the noise). In the imaging mode, {10um resolution can be obtained,where you image a chosen photoelectron energy, so the distribution of a particular chemical shift can be seen, at that sort of resolution. Newer instruments canimage down to 1
-2um. Nevertheless, *if* you can obtain the data you want on an
sem or probe, then maybe, and I don't really have any experience of looking for
chemical shift in wdx or edx spectra, you can more easily do so in small regionsof the sample, and perhaps with shorter acquisition times.

Keith Hallam
Interface Analysis Centre
University of Bristol
England






From: rutledge phil :      prutle1-at-umbc.edu
Date: Thu, 2 Jun 1994 08:24:00 -0400 (EDT)
Subject: Apology to AMRAY

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To all EM users:

Yesterday I sent a message about the Hitachi vs AMRAY problem.
Unfortunately after sending it I read it again and it seemed as if I was
giving a dim opinion of the scopes AMRAY produces. That was not my
intention at all. I got the company mixed up with another company I had
dealings with and with the other company I did have bad experiences with.
It was another American company and I haven't heard anything about that
company in a long time. It could be that company is no longer in
business. I don't know.
To AMRAY I humbly apologize. I know your sales people and have a
deep respect for them. It was totaly unprofessional for me to make a
mistake like this without reading what I sent out.
Again, I humbly apologize.

Phillip Rutledge




From: gkennedy-at-ucsd.edu
Date: Thu, 2 Jun 1994 08:39:31 -0700
Subject: fluorescent dyes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There is a magnificent handbook of fluorescent probes/dyes published by Molecular Probes, Inc. of Eugene, OR.: FAX (503) 344-6504 or telephone (503) 465-8300. It serves as their catalog, and gives really extensive, comprehensive descriptions AND bibliogr
aphies. They do have worldwide distributors, but I don't see any listed for Israel--let me know if I can help any further. Good luck. Grace Kennedy, Scripps Inst. of Oceanography/UCSD (neurobiology, too.)





From: Daniel Henne :      henne-at-sfu.ca
Date: Thu, 2 Jun 1994 11:33:09 -0700 (PDT)
Subject: XEDS;SEM;Standardizing spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello, humble Masters student here.
My question is regarding the standardization of one samples x-ray
spectrum to another. If my SEM was nice and stable I suppose I could
just take a spectrum for a certain period of time. Or if I had some
nice low noise amplifiers I could look at the sample current and
integrate that. However, my SEM isn't very stable and I'm not sure I
want to mess around with the sample current. Has anyone heard of any
other methods or can point to some articles regarding this subject.

Thanks in advance.
Dan Henne
Simon Fraser University
Vancouver, Canada.
email: henne-at-sfu.ca





From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Thu, 2 Jun 1994 15:34:54 -0400 (EDT)
Subject: DAPI

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is in response to the message regarding DAPI nucleic acid stain. The
filter set needed for the DAPI fluorescent stain (bound to DNA) is the UV
excitation filter set; EX max 359 nm and EM max 461.




From: Rodney L Kuehn :      kuehn002-at-maroon.tc.umn.edu
Date: Thu, 2 Jun 1994 15:32:50 -0500 (CDT)
Subject: Re: Amray vs Hitachi

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have a Hitachi SEM and TEM and service is good. The rep is very
responsive to my scheduling problems and works hard to get the scope
running quickly. There is no question about his competence. Neither
scope has ever been down for more than a few days for lack of parts. Yet
I have still not been totally satisfied with Hitachi service because they
depend on the user to know when there is a problem. When Hitachi services
a scope, they fix that problem and leave. They perform no other routine
checks to catch problems early or to deal with problems that might go
unrecognized by a microscopist, particularly if new to EMs. Further, they
have no PM checklist to demonstrate to the microscopist that the scope
really is running at peak performance. In short, the Minnesota rep is
competent and very helpful but is not fully supported by the organization.
AMRAY used to have a marginal reputation but their company seems to have
undergone a complete transformation. I attended one of their user
meetings and it was apparent that their discipline and attention to user
needs was unmatched by other organizations I had examined. I spoke to
several participants and they were extremely enthusiastic about AMRAY
service and many of these people had prior experience with other big-name
manufacturers. For example, after AMRAY services a scope, they do a
routine check on all switches and other items that are reasonably easy and
fast to check. AMRAY also uses an extensive check list on their PM which
is presented to the microscopist along with test photos. I also had the
impression that AMRAY reps are better trained before they are sent into
the field solo. When my rep started here, he didn't know how to stigmate
the TEM!
Of the service meetings I attended (Hitachi, AMRAY, Cambridge, and JEOL),
AMRAY was by far the most aggressive in extracting grievances from their
users. The AMRAY people running the meeting worked their way down a
rather long list, "Has anybody had trouble with response time? Has anyone
had trouble getting parts." They spelled out their standards for response
times so people knew what to expect. I was impressed. If this meeting
(now about 2 years in the past) was typical of the organization, they must
have good scopes and service.
I must note that I have no personal experience on an AMRAY or with their
service. My impressions result from this one AMRAY user meeting.

Rod Kuehn
University of Minnesota







From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Thu, 2 Jun 1994 16:13:29 -0500 (CDT)
Subject: Neutron Scattering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Neutron Scattering/(Diffraction) can be done at the
Intense Pulse Neutron Source (IPNS) Facility
Argonne National Laboratory

Contact the Director: Bruce Brown 708-252-4999
He will fill you in on proposal submission, equipment details
or tell you who to contact....

Nestor Zaluzec
ANL EMCenter




From: troberts-at-eosin.path.uwa.edu.au (Terry Robertson)
Date: Fri, 03 Jun 1994 09:44:46 +0800
Subject: Demonstrating acetyl cholinesterase receptors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Are there any ultrastructural biologists in the email network that have
used any techniques for demonstating acetyl cholinesterase receptors at the
ultrastructural level ? Does anyone know of any source for peroxidase
conjugated, gold conjugated alpha bungarotoxin or an antibody that has been
raised to alpha bungarotoxin?. I would appreciate any information.

Terry Robertson



Dr Terry A. Robertson Telephone: 61-9-3462935
Department of Pathology Fax: 61-9-3462891
University of Western Australia
Nedlands WA 6009
Australia Email:
troberts-at-eosin.path.uwa.edu.au





From: rutledge phil :      prutle1-at-umbc.edu
Date: Fri, 3 Jun 1994 11:30:25 -0400 (EDT)
Subject: Costa Rica/Australia

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


After 2 years of debating about this I decided to start seeking a position
out of the U.S.A. The areas I narrowed it down to are Costa Rica or
Australia. I was wondering if anybody knew of any universities or
private companies in these areas that might have need of an electron
microscopist with 27 years experience. Actually, somewhere in Europe I
would consider also, preferably Sweden, Germany or Switzerland.
Thanks,

Phil Rutledge




From: Susan Smith - National Steel Corporation :      smiths-at-mlc.lib.mi.us
Date: Fri, 3 Jun 1994 11:46:59 -0400 (EDT)
Subject: UV Wavelengths

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a book that lists emission wavelengths for the
elements in the 1000 to 2000 angstrom range? We can find some info where
the wavelengths are listed by element, but we would like a listing of WLs
in order of WLs, in order to check for interferences.

TIA,

Sue Smith
National Steel Corp.
smiths-at-mlc.lib.mi.us




From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Fri, 3 Jun 1994 13:30:45 -0400 (EDT)
Subject: EM jobs/Australia

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Phil,

Try and get in touch with Prof. Brian Gunning, Research School of
Biological Sciences, Australian National University, P.O. Box 475, Canberra
ACT 2601. He may able to suggest the names of persons who may know about
openings for electron microscopists.

Partha




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 3 Jun 1994 13:44:05 -0500 (CDT)
Subject: No Job Hunters Please!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Can I please once again remind you that the listserver is not
to be used by Job hunters looking for positions. I am sensitive
to this issue as I have turned down many requests from both
ANL personnel as well as outside users all qualified and many
being personal friends, all who have ASKED to post this type
of information and were turned down. Given that I just posted
a notice reminding people of this only a week ago I am extremely
disappointed in the recent postings. Once again, you are welcome
to use the MSA/EMC BBS for these purposes, but NOT the Microscopy
Listserver.

If you are not sure about a posting, I would be glad to have
a quick look at it first before forwarding it to the listserver
especially, if you are not sure whether or not it is appropriate. I have
already done this many time for subscribers (most of which
were by the way posted without modification). BUT,
I donot want to convert this to a moderated, subscription only
list as I really don't have the time. So please remember or
refresh you memory on the "ground rules" you were sent when
you first subscribed. I didn't ask for much, and your getting
alot for almost nothing. I will also consider the suggestion
that was sent to me by several subscribers of removing repeat
"offenders" from the listserver, if someone gets abusive.

:-( Nestor Zaluzec - ANL EMCenter





From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Fri, 3 Jun 1994 12:36:18 -0500
Subject: Re: XEDS;SEM;Standardizing spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dan Henne asked about standardizing beam current for taking x-ray spectra:

First of all I agree with what Bob Craig said in his nice n' concise summary of
the do's and don't's of collecting "calibrated" x-ray spectra. I'll specifically
address Dan's question of how to set up a calibrated, reproducible beam current
with out using any kind of current meter. Of course, you won't know the *value*
of the beam current, but at least you'll know it is the *same* value from sample
to sample and day to day.

We do not have a specimen current meter nor a Faraday incident beam current
meter, so what we do in my lab to standardize beam current is the following:
Mount a small metal foil, like aluminum or copper, onto the edge of each stub
that the samples are mounted on using carbon paint or double stick tape. Run a
little collodal graphite around an edge or two to provide additional adhesive
support and a conductive bridge to the foil. Then before taking spectra, move
the electron beam over to the foil and look at the metal foil count rate. Then
adjust the beam current using the condensor lens, or spot size, control to set
an arbitrary rate, say 3,000 counts per second, near the high end of x-ray
through-put for your system (especially for biological samples; may not need a
high end setting for materials samples), but not more than, say, 33% deadtime.
This value can be checked for stability periodically and should be checked for
each new stub you put into the SEM. Thus you will be delivering the same
incident beam current to your unknown samples. Of course, there is always some
indeterminate error with any kind of instrumental setting, and the count rate
fluctuates a bit from second to second due to detection and counting statistics
in the electronics and the physics of beam current production. With our set-up,
the fluctuation is no more than 100 counts per 3,000 and over a 100 second
analysis time the fluctuations average out. Also, our SEM a Philips 500X, is
quite stable. Provided the filament doesn't shift, the same 3,000 cps will be
there hours later. We set the rate when we put a sample in, and check it about
every 30 minutes to monitor any changes that may occur. If a change has occured,
we must throw out some data and collect over again.

A few precautions: use only clean foils; actually image the foil surface at
moderate magnification to make sure that you place the beam on a clean area
without microscratches. Also, must set the stage tilt and detector-to-sample
distance to some standard and reproducible setings, etc. as Bob Craig pointed
out.

But if your SEM is as "unstable" as you make it sound, you will have to check
the foil count rate immediately befor and after collecting a spectrum and keep
only those spectra where the rate did not change during the analysis time. If it
does change frequently, you may only be able to do limited relative quantitative
analysis based on ratios.

One comment about the use of specimen current (if you're lucky enough to have a
specimen current meter on your SEM) to set incident beam current. As Bob Craig
said, you can't do it by monitoring the current on the actual specimen, UNLESS
you pick the appropriate specimen; for example, keep a margin of the aluminum
stub clean and free of specimen, paint, tape, scratches, etc, and measure the
"specimen" or absorbed current there. This value is the total incident current
minus the backscattered fraction, and this fraction should stay the same as long
as you use the same metal target, kV, geometry, etc, etc, etc, etc. So you are
sampling the same fraction of the total and by setting the same value of
specimen current, you are then standardizing the incident curret. I would think
that this would be another way, like my above mentioned indirect technique based
on foil x-ray production, to consistently set a constant incident beam current.



--

Gib Ahlstrand
Electron Optical Facility
University of Minnesota
Dept. Plant Pathology
495 Borlaug Hall
St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX
giba-at-puccini.crl.umn.edu






From: EMLAB-at-opus.mco.edu
Date: Fri, 03 Jun 1994 16:10:36 -0400 (EDT)
Subject: Re: Backscatter detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are in the market for a backscatter detector for our Cambridge 180 SEM.
I ask your opinions on makes,models, etc. for such an item. We mainly
are a life science (pathology) lab thus any info relating to this area is
appreciated.
Also any info, opinions, etc. on multi-channel-plate detectors vs backscatter
detector? Are they worth the extra dollars? (for a life science lab)

Thanks in advance for your 2 cents worth.

Ed Calomeni





From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Fri, 3 Jun 1994 17:39:59 -0400 (EDT)
Subject: Protein Crystals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Gary,

I was not sure wheather you are looking at plant cells or animal cells.
If it is plant cells, there several publications. But if you want
a single source perhaps the one by Behnke would be most helpful:

Behnke, H-D. (1991). Nondispersive protein bodies in sieve elements.
A survey and review of their origin, distribution and taxonomic
significance. Internatl. Assoc. Wood Anat. (IAWA) Bulletin 12:143-175.

Partha




From: jacobb-at-ux5.lbl.gov
Date: Fri, 3 Jun 1994 17:29:33 -0800
Subject: Re: AMRAY

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199406040021.RAA28735-at-ux5.lbl.gov}

We have been dealing with AMRAY for 17 years now. There service has been
outstanding here in the Bay Area: they respond quickly, home in on the
problem, and do what needs to be done. They do routine maintenance. Our
1000A microscope is getting along now but it has always made very good
pictures and remains an excellent low-resolution instrument: it runs and
runs with little down time.

Jacob Bastacky
Jacob Bastacky, MD
Lawrence Berkeley Laboratory EMail: sjbastacky-at-lbl.gov
University of California Phone: (510) 486-4606
Berkeley, California 94720 Fax: (510) 486-4750





From: {henne-at-sfu.ca}:ddn:wpafb
Date: 6-2-94 2:34pm
Subject: Re: XEDS;SEM;Standardizing spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9406040024.AA03054-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: XEDS;SEM;Standardizing spectrum
Orig-Author: {Daniel Henne {henne-at-sfu.ca} }:ddn:wpafb
-----------------------------------------------------------
Hello, humble Masters student here.
My question is regarding the standardization of one samples x-ray
spectrum to another. If my SEM was nice and stable I suppose I could
just take a spectrum for a certain period of time. Or if I had some
nice low noise amplifiers I could look at the sample current and
integrate that. However, my SEM isn't very stable and I'm not sure I
want to mess around with the sample current. Has anyone heard of any
other methods or can point to some articles regarding this subject.

Thanks in advance.
Dan Henne
Simon Fraser University
Vancouver, Canada.
email: henne-at-sfu.ca







From: PHMOULDK-at-usthk.ust.hk
Date: Sat, 4 Jun 1994 11:47:18 HKT
Subject: Re: Ion Tech Ltd.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ion Tech Ltd was taken over a few years ago - I cannot remember by who.
However, they are still in business but trading under the name of Atom Tech
Ltd.

Atom Tech Ltd can be reached in the UK at:

Island Farm Avenue,
West Molesey,
Surry KT8 2UZ,
England.

Tel. 44 (0) 81-941-8959
Fax 44 (0) 81-941-8948

I hope this helps.

Keith Moulding.




From: KOLAR-at-ASUHRM.LA.ASU.EDU (HARRY R. KOLAR)
Date: Sat, 4 Jun 1994 15:29:41 -0700 (MST)
Subject: Looking for conference info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Full-Name:
Reply-To: minter-at-Kodak.COM


I have been unsuccessful in locating ANY information on the International
Symposium on the Structure and Properties of Dislocations in Semiconductors to
be held in 1995 (I believe in Italy). Would anyone happen to have anything
on this conference or direct me to someone who does? Any assistance would
very much be appreciated. Thank you.

H. R. Kolar
Center for Solid State Science
Arizona State University
FAX: (602) 965-9004




From: XLWANG-at-utkvx.utk.edu
Date: Sun, 05 Jun 1994 12:31:15 -0400 (EDT)
Subject: SUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


SUBSCRIBE MICROSCOPY-at-ANLEMC.MSD.ANL.GOV




From: XLWANG-at-utkvx.utk.edu
Date: Sun, 05 Jun 1994 12:31:15 -0400 (EDT)
Subject: SUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


SUBSCRIBE MICROSCOPY-at-ANLEMC.MSD.ANL.GOV




From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Mon, 6 Jun 1994 09:29:02 -0400 (EDT)
Subject: Fee for service facilities, request for info on administration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Apologies to anybody who received this twice due to cross posting.

Every once in a while somebody asks for a survey of user fees for
multi-user facilities involving LM or EM.

The discussion of fees has been very informative.

We were wondering, in addition to fees, how these facilities are
administrated.

For instance:
o What is the administrative hierarchy?
o What is the salary and rank of the director of day-to-day operations?
o What are the responsibilities of this person or people?
o Who contributes the equipment to the facility and how are these
resources allocated to users?

Any comments would be appreciated.

Thank you.

Michael Cammer
cammer-at-aecom.yu.edu
cammer1-at-telico.bioc.aecom.yu.edu






From: ARGIL-at-delphi.com
Date: Fri, 03 Jun 1994 21:00:49 -0400 (EDT)
Subject: Film holder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have a microscope that came with a camera adapter that, among
other things, has a film holder for 2 3/4 x 3 1/2 sheet film. Has
anyone ever used film of this size? I would certainly appreciate any
information that I could gather. I am not familiar with this size, or
why it would be used.

Thanks,

Arthur Gillman
Princeton, NJ





From: tivol-at-tethys.ph.albany.edu
Date: Mon, 06 Jun 1994 10:52:45 EDT
Subject: 6.5 * 9 cm Film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Arthur,

At the HVEM we use 6.5 * 9 cm film (~= 2 3/4 x 3 1/2). This film is
satisfactory for our purposes and the larger standard size will not fit in the
camera. Either Kodak 4489--a very fine grain film which is our usual choice--
or SO163 can be obtained in this size. We also have some old x-ray film cut to
this size which is used primarily for recording diffraction data. If you de-
cide you don't want to get into this size film, could you please send me a de-
scription of the holders you have? By some miracle, they may be compatable
with the AEI holders we use. Thanx.

Yours,

Bill Tivol




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 6 Jun 1994 11:14:29 -0500 (CDT)
Subject: Monte Carlo

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John


The EMMPDL has a contributed MC program called
MacSim. Here is the abstract...

Nestor

Title :MONTE-CARLO SIMULATION (Z's MC v. 1.1)
Keywords :.Electron-beam, Monte-Carlo simulation, energy loss,
backscattered electrons, Rutherford cross section, Mott cross section.
Computer :.Macintosh
Operating System :6.0 or 7.0
Programming Language :Pascal (Think Pascal)
Hardware Requirements :standard
Author(s) :.Zbigniew J. Radzimski.
Correspondence Address :.Analytical Instrumentation Facility
North Carolina State University
Raleigh, NC 27695-7916
Abstract:
The current program is the further development of the program written many years
ago by R. Myklebust and D.C. Joy and modified later by J. C. Russ (J. of
Computer Assisted Microscopy, 1990, 2(2), p. 59). It calculates various
parameters related to electron-beam interactions with solids related to absorbed
and backscatetred electrons. It accepts multi-element and multi-component
structures. It uses Rutherford or Mott cross section for scattering. Special
thanks to Zbigniew Czyzewski and David Joy for providing Mott tables (see J.
Appl. Phys., 1990, 68(7), p. 3066).





From: Larry Hawkey :      hawkey-at-neuro.duke.edu
Date: Mon, 6 Jun 1994 11:15:24 -0400
Subject: EM training

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I know someone who is helping to set up an EM lab. They are interested
in finding out where and how they might send someone to get training.
I know that the various makers of EMs offer courses on how to use their
instruments, but where can they get training on processing, sectioning,
and other techniques used for boilogical TEM. (used for examining
ultrastructure). Any information will be helpful.

Larry Hawkey
hawkey-at-neuro.duke.edu




From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Mon, 6 Jun 1994 13:33:12 -0400 (EDT)
Subject: Courses in Biol. EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Larry,

There are several universities that offer courses in the techniques
of biological electron microscopy including Cornell. Some years ago, I
think the Microscopy Society of America (MSA) sponsored a survey of EM courses
offered in the U.S. universities. If you get in touch with MSA they may be
able to send you a copy of the survey.

Partha
M.V. Parthasarathy
Section of Plant Biology
Cornell University, Ithaca, NY 14853. USA
Tel: 607-255-1734
Fax: 607-255-5407




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Mon, 6 Jun 1994 16:52:26 -0400 (EDT)
Subject: Re: Fee for service facilities, request for info on administration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Our facility is designed to be totally self
funded. Administration heirarchy is:

Faculty director (25% time)
Sets goals, policy, etc.
Consults with users and helps design projects
Sees to overall day to day administration

Facilities engineer (full time)
Maintains microscopes (we 2 SEMs, TEM, IVEM)
Trains users
Developes film
Helps with specialized procedures

2 EM technicians (full time, each)
Processing and embedding
Specialized procedures
Immunocytochemistry
Enzyme cytochemistry
Autoradiography
Etc.
Sectioning
Printing
Maintain EM records
Billing






From: Susan Smith - National Steel Corporation :      smiths-at-mlc.lib.mi.us
Date: Mon, 6 Jun 1994 17:03:05 -0400 (EDT)
Subject: Re: Film holder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Contact Kodak in Rochester, NY. Ask for the Scientific Advisor/Advice
Department.

Sue Smith for Sam Purdy -at- National Steel, Trenton, MI

On Fri, 3 Jun 1994 ARGIL-at-delphi.com wrote:

} I have a microscope that came with a camera adapter that, among
} other things, has a film holder for 2 3/4 x 3 1/2 sheet film. Has
} anyone ever used film of this size? I would certainly appreciate any
} information that I could gather. I am not familiar with this size, or
} why it would be used.
}
} Thanks,
}
} Arthur Gillman
} Princeton, NJ
}
}




From: JJMILL-at-bunyip.ph.rmit.oz.au
Date: Tue, 7 Jun 1994 08:40:03 EST-10
Subject: user fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.940607084003.448-at-bunyip.ph.rmit.edu.au}
To: microscopy-at-anlemc.msd.anl.gov

Re : Fees for use of facilities in Em Units

Since we have only recently joined this group, we missed the
comments on user fees which have apparently been circulating from
time to time. Does anyone have a summary ?
We are interested in not only fees for individuals or groups external
to the university, but also we are obliged to cross charge for all
teaching and research within the university to support the Unit.
What scale of fees are commonly used for SEM, TEM,EPMA, spec
prep from simple to complex ?
Are there "discounts" for local users or are there different rates ?
Are the rates different for expert users ?
etc, etc....
THank you in advance for the comments.

John Millar
EM Unit, RMIT, Melbourne Australia
jjmill-at-rmit.edu.au




From: Weislaw Jablonski :      wis-at-lab.csl.utas.edu.au
Date: Tue, 7 Jun 1994 10:22:53 +1000
Subject: message for Ed Calomeni about BSE detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There are at least 2 very good BSE detectors assemblies namely Robinson BSE
and Oxford Instruments P/L TETRA BSE. Robinson BSE is well known for its
performance including biological subject but I was most impressed by per
formance of TETRA BSE. It is auto-biased with four independent quadrants,
which can give variety of BSE signal combination and working with a electron
beam treshold of 0.1 nanoamp--I think-- it is a jewel. Have seen it working
at ElectroScan Corp- Boston installed with E-SEM. Regards and good luck with
the purchase , Wis Jablonski CSL Uni Tas ,Tasmania, Australia




From: Dwight Beebe :      beebed-at-ERE.UMontreal.CA
Date: Mon, 6 Jun 1994 23:15:45 -0400 (EDT)
Subject: EM - Facility size and # of users?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
These are follow-up questions for Jay Jerome's posting regarding
his unit: how many people does your lab serve? What percentage are from
"in-house" (your department/division) vs. those from "outside"? Do you
charge different fees based on the affiliation of the user and their
level of expertise? How do you control the "quality" of the user?
I have worked in the Cornell TEM Facility run by M.V.
Parthasarathy, which seems somewhat similar to yours (with the exception
of no in-house technical support for the instruments), where a user could
pay a flat fee for a certain period, plus extra costs for special
equipment, e.g., high-pressure cryofixation. A user could also pay on a
per hour, plus materials basis. Do you have such a scheme as well?
My department, which has a simple facility with an SEM and a TEM,
charges virtually no fees to department users, except for the materials
(films, stuff used in processing material, etc.) and we are evaluating
the idea of initiating additional user fees.
I realize that this thread has been carried for quite a while in
this group, so thanks for the patience. Because this could be rather a
contentious proposal for the department users, I'd like to hear from as
many of you who have or had such user fee programs in place.
Thanks very much!

Dwight U. Beebe
Institut de recherche en biologie vegetale beebed-at-ere.umontreal.ca
Universite de Montreal Voice:514-872-4563
4101, rue Sherbrooke est FAX:514-872-9406
Montreal, PQ H1X 2B2 Canada







From: Marcelle A Gillott :      magem-at-csd4.csd.uwm.edu
Date: Tue, 7 Jun 1994 10:52:28 -0500 (CDT)
Subject: Re: user fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In regard to general info on user fees - I think that there was a survey a
few years ago through MSA (maybe the Education Committee or Tech Forum?)

depending on the size of your operation, and whether you are the only EM
facility on campus, you can charge for everything (which can make people
feel like you ar nickel&diming them to death) or roll the cost of coaters
into the SEM beam time, etc; a two tiered scope charge for assisted vs
unassisted usage is an incentive for folks to learn how to operate the
beast - but be sure you retain the right to decide who is qualified for
unassisted usage; a third level, where you are performing the work might
be the same as the assisted usage charge or different, depending on your
philosophy

***************!!!!!!!!!!!!**********

An important point to ponder in regard to the question of different rates
for different users - be very careful ....... most of us are on federal
funds and many of the instruments were purchased on federal funds -
therefore there are some "rules" that have to be followed - these are the
ones I am aware of - there are probably others (if anyone knows where
these are concisely enumerated I would like that info)

- If the equip was purchased through government funding 0r is supported
by govt funds and you are offering services to the "general public"
(industry, hospitals etc), you cannot undercut the prices of any
locally available commercial service- check the yellow pages for labs in
your area (clinical & industria testing - few of these folks are in MSA)


- Anyone who's research is funded through the feds (don't know if this
also applies to state grants) has to be charged the same price - there
are ways to give the folks in your dept a break if the dept is providing
financial support to the facility, but be sure it is clearly spelled out

- An additional charge = the % overhead your institution takes from fed
research grants may be added to the fee for academics from other
institutions who are grant funded

- there are also restrictions on using depreciation as a charge basis if
the instrument was bought on govt funds and the users are govt funded


- i am sure there are more rules than this but these are the ones I have
stumbled across


maybe one of the msa committees should undertake an investigation into
this question as we all seem to stumbling around in the dark - to my
knowledge no one has yet gotten in hot water on an audit, but as more and
more of us are going (at least in part) to a charge basis, it might be a
good idea to have some guidelines


hope this is helpful

Marcelle A Gillott







From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Tue, 7 Jun 1994 13:07:35 -0400 (EDT)
Subject: Re: EM - Facility size and # of users?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In answer to Dwight Beebe's additional questions:

1. Our laboratory servers approximately 80 major users (repeat more than
2X a year). 9 from our department, 41 from other departments within our
institution, 24 from outside our institution (4 from industry), and 6
from foreign countries.

2. We charge less for departmental users because our department pays part
of the service contract on the instruments.

3. We do not charge less based on user expertise. We provide initial
training in any aspect of microscopy the user desires. After that we
charge a fee for technical time if one of our staff is involved in the
microscopy.

4. As outlined in a previous communique (Nestor- do we have a FAQ
section?), we breakdown our fees for each individual technique done and
cKcharge what it costs us to complete the task.Determining these fees is
a major headache, but simplifies administration later and avoids hurt
feelings since all are charged based only on what they use. We of course
do not nit pick (i.e. charge for each grid, each cc of osmium, etc).
Rather we have an average cost for negative stain, cost for
cryosectioning, cost for cyroEM, etc.

5. Microscope usage is on a sign up reservation basis. Inexperienced
users can only sign up during hours when resource staff are available for
consultation. Once we (as a group) are convinced a user can handle
themselves we make the microscopes available to them 24 hours a day.
Rarely, but it has happened, we encounter someone who is either incapable
or unwilling to use the laboratory in a responsible way. These people are
denied all access.

6. Marcelle Gillot's summary of the problems of determining users fees is
excellent and is almost identical to the constraints we use for
determining fees. However, our legal department has a slightly different
interpretation of allowable charges for outside, non-grant supported
work. They advise that it is allowable to undercut the cost of locally
available commercial services as long as you can document (and that may
be a problem) that you do so while still recouping all costs of the work
without relying on subsidies (goverment grants, institutional support for
service contracts, salaries, building maintenance and rent etc.) to cut
your cost. In other words, if
you are willing to take less profit margin or pay your labor sweat shop
wages you can charge less than your commercial competition. You just need
to prove that you and your competition are playing on a level playing
field.

I hope this is helpful.
I would support a movement within MSA to survey and document the issues
involved with fee for service operation. Business is clearly something
most of us where not traine in and really don't want to spend our time
on- however the world being what it is today..........
Regards
Jay Jerome




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 7 Jun 1994 16:59:27 -0500 (CDT)
Subject: SEM:Black Backgrounds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jim Romanow asks about techniques to produce "black backgrounds"

To possibly state the obvious, you could digitize the images (say
using a standard flat bed image scanner ~$1500)
apply a digital mask to the area you want to preserve and then
"Paint" the background black. That technique is used by commerical
artists all the time, many of which are going digital.
I know of some "Electron Microscopy Artist's" who basically
do the same thing using for example Adobe Photoshop on a MacIntosh.

If you want to go analog, then let me assume
that you have tried all the usual tricks like black level shifting
gamma processing and other various electronic signal manipulations
and they don't get you the image contrast you want.

Have you tried exploiting the directional nature of Backscattered
Electron Imaging? Here's what I'd try if I were really pressed
to make the background signal as low as possible.

1.) Use a Low Z stub/base this minimizes the BSE signal
2.) Mount the sample on a fine wire stand.
3.) Tilt the stub/holder so that it is tilted away from
the BSE detector
4.) Bend the support wire so that the sample faces the
BSE detector and so that the sample is between
the wire stand and the detector.
5.) Turn off the secondary electron collection field
and put a negative bias on the collection grid.

Using this configuration, most secondary electrons from the "background" stub
will be rejected by the BSE detector field (I'm of course thinking
of a standard PMT detector with a negative retarding field used to
reject low energy secondaries) and the BSE from the stub/holder will
be for the most part going in the "wrong" direction hence will contribute
neglegibly to the signal. You should get a very dark background signal
here, however, all edges etc will be enhanced on your specimen and your
image may not show the features that you are more comfortable seeing using
conventional secondary imaging.

You could of course get fancy with more digital technology and some
image processing alogrithms. For example, if you want to bring back
the average signal level, record a normal secondary electron image and
:-) digitally add/merge the BSE and SEI signals together on your computer,
to restore some of the lost SEI information.

If the Coates and Welter does produce true NTSC TV rate signal you
could take the signal directly into a TV rate frame grabber board
on a computer and the image process to your hearts content. The
problem will likely be that the "TV-rate" image that your SEM generates
is probably not NTSC but just a fast rate which appears TV like.
Slow scan digital imaging is of course do able, however, it's expensive.
but you said that you didn't have a slow scan mode, so that option
is not viable for you.

I'd probably go the image masking route using Photoshop. If you have
patience then you can get excellent results, but so do those guys with the
scissors and paste!

Nestor J. Zaluzec
ANL EMCenter




From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Tue, 7 Jun 1994 17:29:44 -0500
Subject: EDS Detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lets talk EDS Detectors!
Within a few months time, we will be adding an EDS detector with a thin
window for light element detection capability to an existing TEM (Philips CM12,
LaB6 electron source, no scanning coils (STEM), + or - 60 degree tilt stage,
double tilt analytical grid holder). We are looking at a variety of detector
characteristics and are considering:
1. resolution, 2. crystal area, 3. liquid nitrogen cooled vs. peltier cooled
(dry), 4. silicon vs. germanium detector crystals, and 5. thin window
characteristics. These aspects of detectors are somewhat linked together so its
a matter of determining needs and priorities and convolving them with the
available budget. We have 16 years of experience running EDS (7.0 micron
beryllium window, also soon to be replaced with a new thin window detector) on
an SEM, but the addition of EDS to a TEM, with light element detection, will be
new territory for us.
We are based in a college of agriculture and study a wide variety of
biological samples, soils and minerals, and occasional metallic inclusions.
We have been discussing these issues with vendors of EDS equipment to
enable us to become familiar with the current product line. Now we want to put
some of our reasoning and musings in the above areas to the network for feedback
from our colleagues. We hope that some experienced hands in the EDS buisness
will take a few minutes to respond with the kinds of helpful comments that only
someone who has done real, practical work can give. We apologize for the length
of this communication, but with the advent of Peltier cooled detectors and the
recent introduction of germanium crystals, there is plenty more to consider when
buying a new EDS system these days.

1. Resolution. After years of looking at flabby peaks on our 16 year old EDS
system (resolution about 162 eV -at- Mn Ka), naturally our inclination is to go for
all the resolution we can afford. This seems to imply using a LN cooled premium
quality crystal with a 10mm area. Considering the sensitivity to low energy
x-rays (0.1 to 1.0 KeV) possible with thin windows, it seems one would
particularly want very good resolution to help sort out K,L, M shell overlaps in
this low energy region. For light element (or low x-ray energy) detection, a
seperate resolution at Fluorine (e.g., 68 eV -at-F) is usually given and we've seen
values from 65 to 115 eV -at-F in currently available thin window detectors. Some
say "not to worry" about resolution so much because todays peak deconvolution
software will sort it out for you, but is it really that good in this energy
range?

2. Crystal area. We are aware that the prevailing wisdom suggests that a 30mm
(mm squared) crystal should be used on a TEM (with some loss of resolution?) for
the larger solid angle of x-ray collection that they have; necessary for thin
samples which generally yield lower count rates that a bulk sample in an SEM
would give. We wonder if the LaB6 electron source in our TEM will provide enough
beam current to get the count rate up (without frying the sample!) to a
reasonable level, meaning maximum accumulation times of usable spectra in, say,
200 seconds? Also, do 10mm crystals come mounted in a smaller diameter "snout"
which can be moved closer to the sample to increase collection solid angle?

3. LN cooled vs. Peltier cooled ("dry"). (There was a question on the net in
late April about "water cooled" detectors but only one response was offered
which explained that these detectors are Peltier cooled and water is simply used
to carry away heat from the Peltier device). Tho generally a bit more expensive
to aquire initially, the main argument for purchasing a Peltier cooled detector
given by vendors (Noran, Kevex/Fisons) is the economy of operating them; no
hassle with liquid nitrogen to fill twice a week, so saves time and money. Well,
we're running a 10 liter dewar on our older EDS right now, and we use LN for
cold traps and freezing procedures almost daily. So another 15-20 liters per
week(estimate includes loss due to cooling down transfer dewar) isn't going to
break the budget. In fact, I would think an EDS dewar is a better place to store
LN than the industrial-grade 160 liter dewar that it comes in which vents off
more gas than a pack of ardvarks on steroids. Vendor economy calculations
apparently assume that you are using LN exclusively for your EDS dewar so that
LN cost, delivery fees, tank rental, and LN technician salary, etc, are not
spread out over other equipment that consumes LN, so its a maximum LN expense
estimate. The down side seems to be higher initial cost and less resolution tho
some figures we've seen for some 10mm Peltier cooled crystals, however, range
between 138-149 eV (silicon crystals, Noran, Kevex). Now 138 is not bad, but
such a detector lists for about $25,000, the 149 eV for $18,300. And if you go
to a 30mm Peltier cooled crystal (Noran, generally recommended for TEM
applications, see 2. above) the resolution drops to 155 eV -at- MN (112eV -at- F).
Is anyone out there operating a Peltier cooled detector on TEM or SEM?
They havn't been around as long as LN cooled ones, so we wonder about
performance and reliability over time. They also require a seperate water
cooling and circulating system to remove the heat being pumped out by the
Peltier device. Any problems or commens about that? How's the resolution , etc.?

4. Silicon vs. germanium crystal. The new kid on the block is germanium. And to
look at its resolution and countrate through-put values, one wonders if
silicon's days are numbered. For SEM only, Link-Oxford offers a thin window 10mm
germanium detector with a resolution of 115 eV -at-Mn at 1,000 counts per second
(cps), 65 eV -at-F; it rises to only 133 eV -at-Mn at 10,000 cps, ($22,200). It has
another slightly lower cost 10mm Ge crystal wth 120 eV -at-Mn, 68-70 eV -at-F
($20,800). For TEM, Noran offers a 30mm germanium crystal with 129 eV -at-Mn. These
resolution values look pretty terrific to us!!
Less important to us, but another advantage of Ge crystals, is that they
absorb higher energy x-rays much better than Si crystals. Silicon absorption
efficiency drops rapidly above about 20 KeV (the unabsorbed x-rays pass through
the crystal without interacting); Germanium absorption efficiency doesn't begin
to drop significantly until x-ray energies get above about 60 KeV. So good
detection of high energy K-shell x-rays is available with Ge crystals. But how
about at the very low x-ray energies of the light elements; does germanium do OK
there? Does it have a dead layer like Silicon crystals do? Are there reasons why
germanium detectors with thin windows are not the choice to make for serious
light element analysis?
Other than being a bit pricey, what are the drawbacks, if any, to Ge
crystal detectors?. They havn't been on the market very long, so perhaps there
are the usual problems of maintaining performance and reliability until the bugs
are worked out. Other than that, it seems to have that "wave of the future" look
to us.

5. Thin windows. We read Mark Lund's helpful, but brief, comments on thin
windows in late April and look forward to his upcoming book chapter on the
subject ( For the MAS, edited by D. Williams and D. Newbury). There are thins
and ultra-thins, depending on how sensitive you need to be to the lightest
elements, Boron and Beryllium. We only need to see down to Carbon here so a thin
window is probably best for our needs. We are concerned about thin window
failure rates, especially on the SEM. Do they need to be cleaned often to
maintain good low energy x-ray transmission? What other concerns should we have
about their care and operation?
As Lund pointed out in his comments, if you are specializing in light
element analysis, you really have to look at the over-all EDS system in addition
to the window material. We do not intend to specialize in light element
analysis, but we do want reasonably good detectability, qualitative and
"semi-quantitative" of C, N, and O. But we also need to look a lot at the range
Z=11 to Z=30, and occassionally above. Does one need specialized software for
light element analysis and does it compromise analysis of heavier elements that
may be in the same spectrum?

Enough already! Lets give EDS detectors a bit of network time. I'll compile
responses into a single file and pump it back out in a few weeks. Thanks in
advance to all who respond to this query.





--

Gib Ahlstrand
Electron Optical Facility
University of Minnesota
Dept. Plant Pathology
495 Borlaug Hall
St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX
giba-at-puccini.crl.umn.edu






From: tim-at-phlogiston.nist.gov (Tim Foecke)
Date: Tue, 07 Jun 1994 14:18:02 -0400
Subject: Grey-scale printers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Any input as to good grey-scale printers, as in TRUE
grey scale. I'm interested in one that can be accessed
from the Mac environment, not thermal paper, and makes
nice transparencies. $$ is of minor concern.

I apologize if this repeats a thread.

Tim Foecke, Metallurgy Division, NIST
tfoecke-at-nist.gov





From: randy nessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Wed, 8 Jun 1994 09:50:39 -0500 (CDT)
Subject: Re: EDS Detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


___________________________________________________________________
Randy Nessler
rnessler-at-uiowa.edu
The University of Iowa Central Electron Microscopy Research Facility


On Tue, 7 Jun 1994, Gib Ahlstrand wrote:

} Lets talk EDS Detectors!
} Within a few months time, we will be adding an EDS detector with a thin
} window for light element detection capability to an existing TEM (Philips CM12,
} LaB6 electron source, no scanning coils (STEM), + or - 60 degree tilt stage,
} double tilt analytical grid holder). We are looking at a variety of detector
} characteristics and are considering:

My question is, without STEM, how are you planning to position
your beam on the area of the specimen that you would like elemental
information from? I am embarrassed to ask this question, as I obviously
haven't researched this in publications. When I asked about collecting
x-ray data while performing diffraction, I was told by a service engineer
that it would flood the detector with BSE's. Also, there is the problem
of diffraction flooding a rather large area of the specimen, and the
selected area diffraction aperture allowing the pattern of information to
be viewed, while blocking out extraenous information.
Are you just interested in average area analysis? I hope I am not
the only one curious about this. This is a little off of the topic, and I
hope not to detract from the conversation. We are looking to upgrade a
detector on a SEM, so I am very interested in following this discussion.
Randy








From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 8 Jun 1994 10:32:41 -0500 (CDT)
Subject: Gray Scale Printers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




The ANL EMCenter uses a Tek Phaser IISDX dye sublimination
printer for all high quality gray scale (& color) work from
computers. It is on an Ethernet backbone and directly
usable from your Mac "chooser". The quality and price
performance for both glossy prints & transparencies is
excellent. Cost ~ $10K for the unit then the price/print
for 8x10" $2 B&W, $3 Color add an extra $1 to each
if you want a transparency.

I'll bring examples to the computer workshop at this year's
meeting in New Orleans if anyone want's to have a closer look
at real EM data versus manufacturer's demo prints.


Nestor Zaluzec
ANL EMCenter




From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Wed, 8 Jun 1994 12:58:00 -0500
Subject: SEM black backgrounds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jim Romanow asked about how to get black backgrounds in SEM images:

I've accomplished that using a few simple specimen preparation techniques. The
basic trick is to produce a very smooth featureless background, low Z
composition prefered:

1. I've done the following for pollen grains or rust spores (dried down onto
glass coverslip support surface from water or ethanol, they just stick
naturally), should work with insects too, provided that any paint you use to
mount the insect to the glass coverslip does not spread out beyond the edges of
the insect so you don't see it in the background. Your pin mount technique
should work here too: Mount a cleaned (EtOH rinse, air dried) 1 cm diameter
round glass coverslip onto a standard aluminum stub using double stick tape and
edge the coverslip with carbon paint to provide a conductive path to the
aluminum and for extra adhesion. Mount your insect to the coverslip using a tiny
dot of carbon paint, or using your pin mount method. Heavy metal coat on a
rotary stage as usual and view in the SEM. In spite of the fact that the
glass(silicon & oxygen) coverslip is dense and coated with gold-palladium, the
pollen grains are brighter than the background and when printing the neg its
easy to drop the featureless background into the black, perhaps with a bit of
doging of the subject, if it has a simple profile, to burn in the background.
The key here is that the background is featureless so that there are no edges of
paint globs or other debris which will light up in secondary electron imaging.

2. Similar to the above, is to use a smooth surfaced double stick tape to mount
your samples to avoid using paint (for very small samples).Or as a smooth
background for your pin mount technique. Some double stick tape surfaces develop
tiny cracks or hole patterns in them as a result of being in the vacuum or
whatever and may not provide a feature-free background. Spreading carbon paint
with a brush to get a low Z background may work if the paint is quite thin but
sometimes it leaves too many "brushstrokes" visible and that detail may show up
in the background.

3. Another method to get a fairly good low-Z background with a high-Z coating on
your sample: Mount samples (directly or pin-mounted) on double stick tape or
with a spot of carbon paint onto a stub that has been previously painted with
carbon paint (and dried) to hide the aluminum, so you have a low-Z background.
Set up your vacuum evaporator rotary stage to zero degrees tilt if you can:
We have an old Ladd rotary tilt stage whose "whirling disk" can be re-mounted to
zero degrees tilt, that is, horizontal. Set up the evaporator for an overhead
carbon rod evaporation on one set of electrodes; set up another set of
electrodes for a low angle metal (Au-Pd) coating of about 7 to 10 degrees
elevation. After pump-down, set the stage rotating and do the two coatings, in
either order.
Very little metal coating will hit the stub surface because of the low angle
of coating. A lot of metal will hit the target, tho it may be a bit thin on top
of the target, hence the overhead carbon coating to get some extra conductivity
on top surfaces. This method has worked very well for imaging seeds from 0.05 to
0.4 mm diameter in backscattered imaging (I wanted even "overhead" illumination
that BSE gives) with black backgrounds. Should also work OK for secondary
electron imaging (gives typical "side" illumination).

Good luck!!











--

Gib Ahlstrand
Electron Optical Facility
University of Minnesota
Dept. Plant Pathology
495 Borlaug Hall
St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX
giba-at-puccini.crl.umn.edu






From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 8 Jun 1994 14:51:01 -0500 (CDT)
Subject: Round Robin Printing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Printer Testing::::::

Stuart McKernan suggested that a set of images be
organized for testing of printers. The results for
which could be put on display at the Computer workshop
at the MSA New Orleans meeting. I'm willing to organize
this if there is enough interest. If you want to participate
Email me seperately (Zaluzec-at-anlemc.msd.anl.gov)

I will collect a set of images from the MSA/EMC
library and then setup a FTPable account
where interested parties can download the files.

My first impression is that we should have the
following in the collection for testing.


SEM-gray scale - Life Science (a bug &/or plant)
SEM-gray scale - Physical Science (a computer chip
&/or fracture surface)
SEM-gray scale - BSE Channeling Pattern

TEM-gray scale - Life Science (cells & collected)
TEM-gray scale - Physical Science (Dislocations & Grain Boundary)
TEM-gray scale - HREM image of a crystal
TEM-gray scale - Electron Diffraction SAED
TEM-gray scale - Electron Diffraction CBED

AEM-Colorized - X-ray MAP and/or EELS Map
LM -Color - ?Suggestion Welcome here



I'm open to more suggestions for test images, make them off line and I'll
collect the responses over the next 2 weeks or so. I'll then post
the results to the mailing list for further comment before we
actually do anything else. I'll attempt to pick micrographs which
push the limit of the printers (max contrast & resolution) rather
than just a pretty picture. I'll also make sure text is present on
each micrograph so that users can also judge the capabilities of
the printer to reproduce high quality text as well as an image.

Nestor Zaluzec
ANL EMCenter




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 8 Jun 1994 15:20:02 -0500 (CDT)
Subject: Listserver Meeting at MSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Chandler said::

} I think it would be worthwhile to set up an informal meeting related to
} this list server. Of course there will be folks coming to the computer
} lab, but do you think there would be a place for a presentation of the
} server and handouts of how to subscribe/unsubscribe, access, etc? This
} might be a chance for exchange of this kind of information, or a way to get
} some of the interested parties together. Just a thought.


There will be a tutorial session at the MSA meeting entitled

" A HitchHiker's Guide to Microscopy & Microanalysis Using
Telecommunications,Email and the Internet"

by "some guy who wears a hat"


At it I'll have a bunch of handouts about the listserver, the EMMPDL, MSA BBS,
a run down of other aspects of Internet (EMail, Telnet, newsgroups, FTP,
Mosaic, ......) and anything else I can fit in given time & space & my
fortitute. We also plan on-line demos of each if the phone lines can
handle the links.

In addition, John Mansfield will be also doing a Tutorial on Image Processing
using NIH Image as an example.

These tutorials are set for Thursday Afternoon August 4th, from 1--} Closing.
(i.e. ~ 3-4 hours), depending on the audience.

Jay Jerome & John Posthill get the credit for twisting John's and my arms into
doing these tutorials. I guess they figure there might be as many as a dozen
people that might be interested, anyway, I know of at least 2 people that
will be there. So Listserver subscribers, you're welcome to stop by, it's as
good a time as any.

The workshop has a reasonably large room again this year, and will be set up
for informal as well as formal presentations throughout the week. Of course, I'm
more than willing to participate in an even more informal get together
at any local pub, so John C. just set a time & place! ;-)

Nestor Zaluzec
ANL EMCenter




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Wed, 8 Jun 1994 17:20:13 -0400 (EDT)
Subject: Re: Listserver Meeting at MSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





}
} The workshop has a reasonably large room again this year, and will be set up
} for informal as well as formal presentations throughout the week. Of course, I'm
} more than willing to participate in an even more informal get together
} at any local pub, so John C. just set a time & place! ;-)
}
} Nestor Zaluzec
} ANL EMCenter
}
FAQ-
Who's buying?
;-}
Jay Jerome






From: Jan Coetzee EM Univ Pretoria :      JANC-at-ccnet.up.ac.za
Date: Thu, 9 Jun 1994 09:46:26 GMT+2
Subject: re: SEM black background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


As various contributors have pointed out, a black background depends on
having the highest possible difference in Z between foreground and a
{featureless} background.

With biological samples it is useful to expose the specimen to Osmium (fix
or postfix in osmium tetroxide for wet specimens, expose dry specimens to
osmium tetroxide vapour in a sealed container for a few hours) and then to
mount on a stub coated with gold size.

Japan Gold Size (the stuff that artists use, obtainable from any art supply
shop, we use Winsor & Newton brand but others would also be ok) is a sticky
liquid, containing mostly linoleic acid polymers. A very thin layer of this
on a stub becomes very tacky within 5 or 10 mins at 50 deg C, depending on
the characteristics of the batch, and can then be used in the same way as
double-sided tape. The thin layer on the stub is made by wiping a slight
smear of the liquid over the face of the stub. This soon polymerizes into a
glassy, featureless, nonsticky, non-outgassing, low Z layer.

The polymerization can be hastened by a further period at 50 deg C.
If your specimens are heat sensitive this further bakeout is not necessary,
the process will go to completion at room temp within a few hours.
Sputtercoat the sample with a minimum of gold and the high Z osmium in
the sample on the smooth low Z gold size will give a nice dark background.
Even a very thin layer of gold size on the stub is thick enough to keep the
beam away from the relatively high Z stub.

The gold size idea is not our own, it was published long ago in a paper
that we cannot now find, but it works really well for a wide variety of
specimens. In addition it is cheap, natural, biodegradeable, nontoxic and
apparently also non-allergenic!
Jan Coetzee
Unit for Electron Microscopy Tel:+27-12-420-2075
University of Pretoria Fax:+27-12-342-1738
Pretoria Internet:janc-at-ccnet.up.ac.za
0002 South Africa




From: rms-at-vax.ox.ac.uk
Date: Thu, 09 Jun 1994 10:02:51 +0100
Subject: Free to a good home!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk

Free to a good home. Philips EM 300 Electron Microscope now surplus to
requirements due to rationalization of Faculty EM Services. Very Good
Condition. Philips maintain the instrument. Further details: Mr J N Brown,
Department of Physiology, University of Edinburgh, Teviot Place, Edinburgh, UK.
Telephone (031) 650 3273, fax (031) 650 6527.





From: pabuffat-at-i2msg1.epfl.ch (Philippe Buffat)
Date: Thu, 9 Jun 1994 10:38:09 +0000
Subject: Free to a good home!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable
To: Microscopy-at-anlemc.msd.anl.gov

=46ollowing the question from R. Nessler:
My question is, without STEM, how are you planning to position
} your beam on the area of the specimen that you would like elemental
} information from? I am embarrassed to ask this question, as I obviously
} haven't researched this in publications. When I asked about collecting
} x-ray data while performing diffraction, I was told by a service engineer
} that it would flood the detector with BSE's. Also, there is the problem
} of diffraction flooding a rather large area of the specimen, and the
} selected area diffraction aperture allowing the pattern of information to
} be viewed, while blocking out extraenous information.
} Are you just interested in average area analysis? I hope I am not
} the only one curious about this. This is a little off of the topic, and I
} hope not to detract from the conversation. We are looking to upgrade a
} detector on a SEM, so I am very interested in following this discussion.
} Randy

In the CM12, you can work under normal TEM observation and reduce step by
step the size of the illumination spot down to about 20 nm while still
retaining enough beam current in the probe to perform EDS microanalysis in
a reasonable time. For that purpose you will use the spot size knob
(physically the 1st condenser lens). This way is available on most modern
TEMs too. DON'T FORGET TO REMOVE THE OBJECTIVE APERTURE, which otherwise
will produce a hudge amount of X-rays and dazzle the EDS detector (it may
take some minutes to fully recover its normal dead time!).
It can be helpful to switch to diffraction observation if you want to
analyze a tiny precipitate that give a different diffraction pattern than
the surroundings. Be careful, a probe shift may happen due to the weak
coupling of magnetic fields of the diffraction lens and the objective lens.
So you will have to slowly shift (blind) the probe until you find again the
caracteristic diffraction pattern of your phase. Then start microanalysis.
If you see the pattern changing during the analysis, you should stop
acqisition, shift the probe again and then continue the analysis.
I desagree entirely with your service engineer. Switching to diffraction
do not change anything to the magnetic field around the sample (ie, the
objective lens) within the limit of the weak coupling mentioned above. So
there are no more BSEs reaching the detector in one mode than in the other.
Thus there is no reason not to use diffraction instead of image if you feel
easier to decide wether or not you are firing on the right phase. He was
probably thinking to the observation in low magnification mode (LM mode for
Philips) where the excitation of objective lens is switched off or very
strongly reduced. Then BSEs do no longer spiral in the magnetic field and
then are allowed to escape toward the EDS detector.

I hope this will help you. Do not hesitate to contact me again if I wasn't
clear. But I am leaving Friday evening for the next two weeks.
Best regards


__________________________________________________________________
Philippe Buffat Ecole Polytechnique Federale de
Lausanne (EPFL)
Institut
Interd=E9partemental de Microscopie Electronique
Address: EPFL-I2M, Batiment MX-C, CH-1015 Lausanne, Switzerland
Phone: +41(21)693 29 83 (Central European Time) Fax: +41(21)693 44 01
______________________ Eudora 2.0.2 __________________________________






From: sking-at-eleceng.ucl.ac.uk (Simon King)
Date: Thu, 9 Jun 1994 13:16:47 +0100
Subject: Re: EDS Detectors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {8136.9406091217-at-busby}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} ___________________________________________________________________
} Randy Nessler
} rnessler-at-uiowa.edu
} The University of Iowa Central Electron Microscopy Research Facility
}
}
} On Tue, 7 Jun 1994, Gib Ahlstrand wrote:
}
} } Lets talk EDS Detectors!
} } Within a few months time, we will be adding an EDS detector with a thin
} } window for light element detection capability to an existing TEM (Philips
} } CM12,
} } LaB6 electron source, no scanning coils (STEM), + or - 60 degree tilt stage,
} } double tilt analytical grid holder). We are looking at a variety of detector
} } characteristics and are considering:

Firstly, thanks for the useful info on EDS detectors.
One question: Has anyone had success with annealing detector crystals to
combat incomplete charge collection?
}
} My question is, without STEM, how are you planning to position
} your beam on the area of the specimen that you would like elemental
} information from? I am embarrassed to ask this question, as I obviously
} haven't researched this in publications. When I asked about collecting
} x-ray data while performing diffraction, I was told by a service engineer
} that it would flood the detector with BSE's.

Switching to diffraction mode focuses the intermediate/diffraction/P1
(nomenclature
depends on manufacturer) lens onto the back focal plane of the objective.
It does (should)
not influence the illumination seen actually at the specimen. Maybe your
service engineer
is thinking of fried (jellied?) EELS detectors ? :-) (which company is he
from, btw?). I've lost the
post now (hereafter referred to as the "lost-poster"), but someone
mentioned weak coupling between INT and OBJ lenses - surely this problem,
re spot shifting, is not a problem with correct rotation centring. Life
starts getting very tricky, however, when your specimen is
charging/magnetic/unstable - I won't go into details here though, since
that's another issue.
For precipitates/inclusions/nano-crystalline materials, viewing the
diffraction pattern is indeed a good way to check the spot position, as
already pointed out. Quantitative microanalysis of these features is a
headache, though, but that's also another thread.

} Also, there is the problem
} of diffraction flooding a rather large area of the specimen,

Use a smaller spotsize (C1 and C2 controls). CBED often is performed with
tiny (~1nm or less FWHM) spot-sizes. LaB6 or FEG will help here. In a
darkened room, you should be able to see sufficient information in the
image formed with the smallest C1 and slightly defocussed C2 to orient the
specimen. Chances are that if you cant see anything, your count rate will
be too low anyway.

and the
} selected area diffraction aperture allowing the pattern of information to
} be viewed, while blocking out extraenous information.

Not sure what you are getting at here. The SAA, being far down the column
(position varies - any post-specimen image plane will do), is well removed
from anything to do with EDS, although possibly it may contribute X-ray
counts if sufficiently high. It's a simple matter to find out if it is
contributing and whether you need to make sure all users remove it prior to
analysis. As the lost-poster said, it is *very* important to remove the Obj
Aperture (OA) or you may start encountering incomplete charge collection
(which manifests itself as an asymmetric, low-energy tail to each peak)
soon afterwards. :-(

} Are you just interested in average area analysis? I hope I am not
} the only one curious about this. This is a little off of the topic, and I
} hope not to detract from the conversation. We are looking to upgrade a
} detector on a SEM, so I am very interested in following this discussion.
} Randy

The comments above refer to TEM/STEM. Life is quite different for SEM.
We too are considering SEM detector upgrade/purchases so are also v
interested in this thread (re detector crystals).






From: streiffe-at-hrem.mpi-stuttgart.mpg.de
Date: Thu, 09 Jun 1994 18:07:49 +0200
Subject: TEM - Dislocation contrast simulation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I'm working on analyzing dislocation arrays in ferroelectric PbTiO3
thin films. I'm searching for any software, preferably for Macintosh or
DEC Alpha UNIX, that will simulate dislocation contrast under various
diffraction conditions. I'm aware of ONEDISANL and TWODISANL (available
through the EMMPDL archive), but would like to try other options if they
are available. Suggestions?

Thanks in advance.

Stephen Streiffer
Max-Planck-Institut f. Metallforschung
Stuttgart, Germany




From: Greg Erdos ICBR EM Core Lab Univers :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Thu, 09 Jun 1994 17:15:25 -0500 (EST)
Subject: Replicas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In the TEM literature there is a technique called pseudo-replica formation
where formvar or collodion is poured in a thin layer over a surface and
alloed to dry. It is then floated off, havin trapped particulates and large
molecules which can then be observed with TEM. Has anyone done or does any
one know of a similar technique for SEM. I need to enumerate particles
attached to moist surfaces that cannot be processed by conventional means for
SEM (living skin). I wonder if pressing scoth tape to the surface would
pick up all the particles down to the several micron range???

Ideas please.

**********************************************************
* Greg Erdos ** *
* Director, ICBR EMCL ** Phone 904-392-1295 *
* 218 Carr Hall ** FAX 904-392-8598 *
* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
* Gainesville, FL 32611 ** *
**********************************************************




From: tvoiles-at-unlinfo.unl.edu (todd voiles)
Date: Thu, 9 Jun 1994 17:05:49 -0500 (CDT)
Subject: Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,


We're seriously considering starting a user-fee type of
arrangement for our electron microscopy facility. I was wondering if
and how much other similar facilities charge for use of microscopes
and equipment as well as consulting, etc..... Our facility is staffed
by a faculty advisor and a half-time facilities manager.

I would even be willing to run a survey and compile the results
if that is what it takes to get this information.

Thanks,

Todd Voiles
Central Facility for Electron Microscopy
Center for Materials Research and Analysis
University of Nebraska, Lincoln, City Campus

tvoiles-at-unlinfo.unl.edu





From: rutledge phil :      prutle1-at-umbc.edu
Date: Fri, 10 Jun 1994 08:15:01 -0400 (EDT)
Subject: Re: Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



On Thu, 9 Jun 1994, todd voiles wrote:

} Hi,
}
}
} We're seriously considering starting a user-fee type of
} arrangement for our electron microscopy facility. I was wondering if
} and how much other similar facilities charge for use of microscopes
} and equipment as well as consulting, etc..... Our facility is staffed
} by a faculty advisor and a half-time facilities manager.
}
} I would even be willing to run a survey and compile the results
} if that is what it takes to get this information.
}
} Thanks,
}
} Todd Voiles
} Central Facility for Electron Microscopy
} Center for Materials Research and Analysis
} University of Nebraska, Lincoln, City Campus
}
} tvoiles-at-unlinfo.unl.edu
} Todd:
Here at UMBC we charge for EM services but since we are a state
institution we can't make a profit. All of our charges cover just the costs.
We are probably on the low end as far as charging for services go. If it
will help, send your fax number and I will fax a copy of our charges to you.
Phil Rutledge
Phone: (410) 455-3582
FAX: (410) 455-3875




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Fri, 10 Jun 1994 12:04:26 -0500 (CDT)
Subject: Nomenclature BBS vs. Listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To: All Microscopy Subscribers:
Re: Administrivia

A minor point, but potentially an important one for the future of
Microscopy Listserver.

The Microscopy Listserver, and the MSA BB system are not the same entity.
Although I basically run both (and also the EMMPDL, & Coordinate the MSA
Computer workshop), the funds for each do not come out of the same pot.

To most of you things are transparent and (relatively?) seemless.
I've tried to keep it that way intentionally as I've tried to make sure that
everyone has access with the maximum amount of versatility (of course
not everything works like it's supposed to but we're getting there slowly).
The reason for this note is that I've seen some references lately
where these admittedly subtle distinctions are accidently blurring, for example,
in references to where things are located or how to get access to some
information. This may not seem like a big deal to most of you, however,
ultimately, I do need to document things to present and future funding sources
(i.e. there is no free lunch) so when and if you talk about these
different systems please try acknowledge them as different beasts.

Just for the record:

The MSA Computer Workshop is funded totally by MSA, except
during joint society meetings, when our sister societies chip
in a proportion of the costs. It is a week long interactive workshop
held each year at the annual meeting of MSA.

The MSA BBS is jointly funded by MSA and ANL EMC. It is a regular
BBS system runing on a PC with both Internet and Telephone links.
MSA supports an 800 number (1-800-MAS-EMSA) as well as having
purchased the PC and some of the software. The ANLEMC pays local costs
(power, phone lines, network lines) and spends time & effort to
keep it updated and running.

The EMMPDL is mainly funded by the ANLEMC, MSA contributes
toward some expenses when local/sister societies want copies, or when
the library is distributed at meetings (for example in the MSA
computer workshop at New Orleans, or local society meetings).

The MICROSCOPY LISTSERVER (i.e. this EMail distributed discussion group)
is run by by the ANLEMC but is not "officially" supported yet by any funding
agency, basically I started it on the side as an experiment (call it a
voluntary service to the microscopy community). Right now the ANL EMCenter
ultimately pays the network costs through it's general overhead costs (so
ANL and DoE should get some credit).If it is referenced, especially
if you want to acknowledge it please simply use the phrase
"The ANLEMC Microscopy Listserver". Try not to confuse
it with the MSA BBS, while I'll know what you mean, if I need to
raise support funds in the future (and I will need to) then it will need to
be acknowledged appropriately.

Those of you who are reading the Microscopy Listserver via Newsgroups
are getting a mail feed from ANLEMC, via a mailserver at the University of
Michigan arranged by John Mansfield. Also,if you are exclusively on a newsgroup
system you should know that the newsgroup replies do not presently filter back
to the Microscopy Listserver subscription list. This means that your replies
and/or questions do not reach most of the users of this system. This is on
the list of things to fix, but it's one of those things that takes time and
$$... If you want to reach the maximum coverage, you should post to:

Microscopy-at-anlemc.msd.anl.gov

Thanks for your patience ....

Nestor J. Zaluzec
ANL EM Center





From: bjfeijo-at-bio.fc.ul.pt (Jose A. Feijo)
Date: Fri, 10 Jun 1994 21:38:47 +0000
Subject: Help on Alden 9315ctp printer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9406092046.AA00266-at-dnagel.bio.fc.ul.pt}
X-Sender: bjfeijo-at-skull.cc.fc.ul.pt
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


We've just received an Alden 9315ctp continuous tone printer, but we were
faced with a surprinsigly poor information for operating the printer,
specially having in mind its price. The only executables they send are a
tprnt.exe, presumably for TIFF files (no indication for the version), and a
couple of conversion executables like t2r.exe and x9pc.exe. Unfortunaly in
our hands these programs seems to have quite an random behaviour, either
printing the image correctly, either producing truncated, layed, or tottaly
corrupted images. So far we've tested a number of programs to generate the
TIf's and Raster's (PSP, XView, PStyler, etc.) and so far we could not find
a pattern for this behaviour. Being in a small european country, the local
rep's can't be of much help (actually it's the first one they sold).
Therefore I would to ask help to anyone who could provide me with any info
on the operation of such printer, including other drivers for direct
printing from the software (Windows for instance), info on net sharing drove
by UNIX machines, and any possible internet contact with the guys from Alden.
Thank you very much for any help.


________________________________________________________________
Jose A. Feijo
Dept. Biologia Vegetal, Fac. Ciencias Lisboa
Ed. C2, Campo Grande, P-1700 LISBOA, PORTUGAL

t. + 351.1.7573141 fax + 351.1.7597716
e.mail bjfeijo-at-bio.fc.ul.pt
URL http://www.fc.ul.pt/departs/biologia_vegetal/index.html
________________________________________________________________





From: Bo Johansen :      BOJ-at-bot.ku.dk
Date: Sun, 12 Jun 1994 09:34:16 GMT+0200
Subject: Anti calmodulin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: Microscopy-at-anlemc.msd.anl.gov

Dear fellow microscopists

I am working on CaM localization and quantification in plant cells,
and for that I have been using polyclonal antibodies raised in rabbit
against bovine CaM. These antibodies were sold by SIGMA, but now
they only sell MAb's, which does not detect the plant CaM.

So is there anybody out there with a stock of polyclonal anti CaM.
It does not matter if the antibodies are raised against plant or animal
CaM, but i would prefere if they were raised in rabbit.

Please answer directly to BoJ-at-bot.ku.dk

Bo Johansen

____________________________________________________________________

Bo Johansen
Botanical Laboratory
Gothersgade 140
DK-1123 Copenhagen K
Denmark

E-Mail: boj-at-bot.ku.dk
Tlf: +45 3532 2150
___________________________________________________________________




From: Marcelle A Gillott :      magem-at-csd4.csd.uwm.edu
Date: Mon, 13 Jun 1994 16:34:50 -0500 (CDT)
Subject: Re: Replicas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Gred Erdos was inquiring about making replicas for SEM of living skin

One method you could try is to use dental molding compound - (the stuff
they use to make impressions for orthodontics etc) - you mix it up, press
it onto the surface of interest and it hardens up very quickly
and can be removed without damaging the original surface -

We have done this a a class exercise - using the replica as a mold to
make a "positive" with Spurrs resin - we generally did 2 applications,
the first being used to remove the small surface debris which you want to
look at so I suspect it should work for you

The stuff we used was called "Cuttersil" and was available through
Columbus Dental
PO Box 620
St Louis Mo 63188

Good Luck

Marcelle Gillott
UWM








From: sbarlow%sunstroke-at-sdsu.edu (Steve Barlow)
Date: Tue, 14 Jun 1994 10:29:21 -0700
Subject: virus on internet

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} } *Received through multiple hands across the sea and
along the internet
} }
} } INTERNET VIRUS ALERT! Date: Wednesday, May 25,
1994
} }
} } A Virus has been discovered on Internet that is
disguised as CD-ROM
shareware.
} }
} } Unknown hackers have illegally put the Chinon name
on a destructive
shareware
} } file and released it on the Internet. This
catastrophic virus is named
} } "CD-IT." --DO NOT DOWNLOAD. IT WILL CORRUPT YOUR
HARD DRIVE. The
program,
} } allegedly a shareware PC utility that will convert
an ordinary CD-ROM
} } drive into a CD-Recordable (CD-R) device, which is
technically
impossible,
} } instead destroys critical system files on a user's
hard drive. The
program
} } also immediately crashes the CPU, forces the user to
reboot, and stays in
} } memory.
} }
} } Widest dissemination is requested.
} }
} }
} }
}
} (Mr) Lindsey Thomas Martin (lmartin-at-sfu.ca)
} Dialogue: Canadian Philosophical Review
} Vancouver Studies in Cognitive Science
} International History Review
} Simon Fraser University, Burnaby BC Canada.
}



----------------------------------------------------------
Dr. Steven Barlow
EM Facility/Biology Dept.
San Diego State University
San Diego CA 92182-0057
phone: (619) 594-4523
fax: (619) 594-5676
email to sbarlow-at-sunstroke.sdsu.edu





From: Paulette Brunner :      pbrunner-at-u.washington.edu
Date: Tue, 14 Jun 1994 11:52:28 -0700 (PDT)
Subject: JEM parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A visiting scholar at the Friday Harbor Labs is looking for parts for a
JEM transmission microscope (model #100B). If anyone knows of a source
please reply to Scott Schwinge at schwinge-at-fhl.washington.edu




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 15 Jun 1994 9:11:44 -0500 (CDT)
Subject: Bouncing Mail Messages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



All Subscribers:

Just so you know, part of the problems which caused a large
number of Undeliverable mail messages has been cured. A server
at the Univ. of Michigan had gone batty and was fixed
by John Mansfield. We still have another server somewhere
in the system which cannot deliver some mail to "previously"
valid addresses. I'm investigating the problem.

Please be patient..... Nestor




From: randy nessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Wed, 15 Jun 1994 10:50:40 -0500 (CDT)
Subject: Re: Non-comos imaging software?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




___________________________________________________________________
Randy Nessler
rnessler-at-uiowa.edu
The University of Iowa Central Electron Microscopy Research Facility


On Wed, 15 Jun 1994, MICHAEL DELANNOY wrote:

} I am looking for software that will drive a confocal microscope but doesn't
} save image files in the .pic form but something more useful for other
} Mac imaging programs such as Photoshop. Alternatively (and more likely) a
} program that converts .pic to .gif or PICT without loss (at least significant
} loss) of image quality. The clumsyness of manipulating images for figures in
} comos is not aceptable
}
}
}
} Shawn Burgess
} c/o Mike Delannoy

Talk to Biorad about the program .pic to .tiff. We have it on our
MRC-600. Lately, people have been copying files to floppies and taking
them to Adobe Photoshop. They are able to manipulate the .pic images knowing
that the header is 76 bytes, there are 768 columns, 512 rows, and
bytes/pixel = 1. Granted, it is a little cumbersome to do this. It
appears that you have the confocal and Mac already, and is proably most
economical to research the means to interface/import images.
Randy







From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 15 Jun 1994 21:55:24 -0500 (CDT)
Subject: Imaging Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Shawn Burgess asked

...... am looking for software that will drive a confocal microscope but doesn't
save image files in the .pic form but something more useful for other
Mac imaging programs ....


You should seriously look into getting a copy of NIH-Image. The current
version level is 1.55 and 1.56 is in beta. You may download a copy
from zippy.nimh.nih.gov using anonymous FTP. Alternatively if you
attend the Microscopy Society of America Meeting in New Orleans in
August you may get a copy at the Computer workshop & software exchange.

Nestor Zaluzec
ANL EMCenter





From: James Drummond :      drummond_james-at-vanlab.paprican.ca
Date: Wed, 15 Jun 1994 15:00:52 PDT
Subject: Any microscopists studying wood morphology?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.940615150052.256-at-vanlab.paprican.ca}
To: Microscopy-at-anlemc.msd.anl.gov

Are there any microscopists out there who've worked on morphology
differences between wood species? We're trying to find data
on northern and interior British Columbia wood species with respect
to fibre dimensions. If anyone has any insight out there I'd be grateful
to hear from you, perhaps directly since this is a topic of extremely
narrow interest and needn't tie up the listserver (unless someone
else wants the info too; let me know). This is also kind of a shot in the
dark to draw some of you forestry related microscopists out of the
woodwork (pun unintentional). Thanks.

James Drummond
Pulp and Paper Research Institute of Canada
Vancouver, B.C.




From: Larry Hawkey :      hawkey-at-neuro.duke.edu
Date: Wed, 15 Jun 1994 16:28:04 -0400
Subject: technical future in EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



First let me say that I AM NOT LOOKING FOR A JOB!
But I would like to see a discussion of the future of technical
jobs in EM. I always look at the classified ads in the EMSA (now MSA)
bulletin. I have always found it interesting to see where tech
jobs were and if my qualifications were keeping me competitive with the
market. The last few MAS bulletin didnot have many, if any, EM
job.

I have had scientist out side EM tell me that EM had had it peak, as
all tecnologies do, and now has less and less uses. I believed
that although pure ultrastructual studies of biological tissue has
been pretty well covered, new advances would renew the need for
trained people.

Is it true that EM is past its prime? EMSA did take out the "E"

Are people advertising job some where else?

Is money so tight that EM has become to expensive to be widely used?

Are service facilities taking the place of numerous smaller labs?

Are old EM tech. so full of Glut and OsO4 that they are not dying off?

Someone please give me some comfort. I still have 28 more years until
retirement.

Larry Hawkey
Hawkey-at-neuro.duke.edu




From: NANCY SMITH :      NSMITH-at-darwin.sci.csuhayward.edu
Date: Wed, 15 Jun 1994 10:49:02 PSD8PDT
Subject: remote control of SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We will be purchasing a pc operated SEM with the intent of being able
to remotely control the instrument. Initially, the remote site will
be in our building and then a second phase is to be able to remotely
control the instrument from a sister campus. Is anyone doing remote
control of a SEM? If so, could you address the following questions?

Have you been able to accomplish remote control other than serial?

What functions can you perform?

What software interfaces have you used to access the SEM from
elsewhere? (carbon copy, pc anywhere, etc).

What are the speed considerations?

Have you attempted to send both video and the data signals in real
time?

How have you accomplished it?

If you have not accomplished this task, do you have any ideas or
suggestions about how it could be done?

Thanks in advance for any help.

NSmith
510-881-3527
FAX 510-727-2035




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Thu, 16 Jun 1994 12:11:54 -0400 (EDT)
Subject: Re: technical future in EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In times of very tight money, I don't think any of us can provide much
comfort. However, I think Larry's points are good ones. I do think more
of us are employing light microscopy (or going back to our original roots
in LM). I think this is good because often LM/EM combined studies can be
much more powerful than single technology. EM provides the resolution,
but LM provides greater sample area and dynamic information. Our scheme
to survive (contact me again in 5 years to see if it works), is to
attract non-traditional EM users (biochemists, physiologists, etc) by
helping design their experiments and making use of EM as easy as
possible. All that is required of them is the ability to focus and
capture the images. This has worked well for us over the years, the
biochemistry department rather than the anatomy or pathology department
constitutes our biggest users. It takes more of my time but it keeps us
solvent. So, yes ithink there is still a large need for us dinosaurs
(molecular genetics still won't answer all questions and a picture is
still worth a thousand words even with inflammation) but you have to be
more of a slalesman and gear your services to the community needs.
My humble opinion.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************


On Wed, 15 Jun 1994, Larry Hawkey wrote:

}
} First let me say that I AM NOT LOOKING FOR A JOB!
} But I would like to see a discussion of the future of technical
} jobs in EM. I always look at the classified ads in the EMSA (now MSA)
} bulletin. I have always found it interesting to see where tech
} jobs were and if my qualifications were keeping me competitive with the
} market. The last few MAS bulletin didnot have many, if any, EM
} job.
}
} I have had scientist out side EM tell me that EM had had it peak, as
} all tecnologies do, and now has less and less uses. I believed
} that although pure ultrastructual studies of biological tissue has
} been pretty well covered, new advances would renew the need for
} trained people.
}
} Is it true that EM is past its prime? EMSA did take out the "E"
}
} Are people advertising job some where else?
}
} Is money so tight that EM has become to expensive to be widely used?
}
} Are service facilities taking the place of numerous smaller labs?
}
} Are old EM tech. so full of Glut and OsO4 that they are not dying off?
}
} Someone please give me some comfort. I still have 28 more years until
} retirement.
}
} Larry Hawkey
} Hawkey-at-neuro.duke.edu
}





From: Greg Erdos ICBR EM Core Lab Univers :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Thu, 16 Jun 1994 12:45:21 -0500 (EST)
Subject: Technical future of EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


EM certainly has a future if our experience is any indication. We run a
central service lab and we have never been without work to do for more than
half a day. Seems just as we catch up (which rarely happens) the phone starts
to ring.

We may not be seeing positions advertised for tech's due to the time
lag involved in the publication of professional society newsletters.
When I have a vacancy in the lab it usually occurs with little more
than a few weeks notice and I need to fill that spot quickly and cannot wait
for the next MSA Bulletin to come out. So I always turn to Jungle Drums for
faster communication.
**********************************************************
* Greg Erdos ** *
* Director, ICBR EMCL ** Phone 904-392-1295 *
* 218 Carr Hall ** FAX 904-392-8598 *
* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
* Gainesville, FL 32611 ** *
**********************************************************




From: A. Kent Christensen :      akc-at-umich.edu
Date: Thu, 16 Jun 1994 13:10:37 -0400 (EDT)
Subject: Re: technical future in EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Wed, 15 Jun 1994, Larry Hawkey wrote:

}
} First let me say that I AM NOT LOOKING FOR A JOB!
} But I would like to see a discussion of the future of technical
} jobs in EM. I always look at the classified ads in the EMSA (now MSA)
} bulletin. I have always found it interesting to see where tech
} jobs were and if my qualifications were keeping me competitive with the
} market. The last few MAS bulletin didnot have many, if any, EM
} job.
}
} I have had scientist out side EM tell me that EM had had it peak, as
} all tecnologies do, and now has less and less uses. I believed
} that although pure ultrastructual studies of biological tissue has
} been pretty well covered, new advances would renew the need for
} trained people.
}
} Is it true that EM is past its prime? EMSA did take out the "E"
}
} Are people advertising job some where else?
}
} Is money so tight that EM has become to expensive to be widely used?
}
} Are service facilities taking the place of numerous smaller labs?
}
} Are old EM tech. so full of Glut and OsO4 that they are not dying off?
}
} Someone please give me some comfort. I still have 28 more years until
} retirement.
}
} Larry Hawkey
} Hawkey-at-neuro.duke.edu
}
Larry,

My own bias is that EM use for biology will increase in the future.
We have all seen a very striking renaissance of morphology at the LM level
over the last few years. For example, everyone in molecular biology now
wants to do immunocytochemistry, in situ hybridization and to localize
reporter gene expression. And molecular biologists are finding out that
they can utilize these methods intelligently only if they have some
understanding of microscopes, tissue organization, cells, ultrastructure
and basic morphological methods and strategies. New technologies are
greatly expanding the capabilities of morphological methods. For example,
the confocal LM can show where fluorescent ICC activity is in cells with
much greater clarity than previous fluorescent microscopes could. All
this is a far cry from a few years ago, when graduate students thought
that if they learned how to clone genes their futures would be assured.

I really believe that these trends will continue, and will extend
more and more to the EM level. It is often important to localize antigens
within cells, which can be done by immunocytochemistry with gold particle
labels. When lacZ (yielding beta galactosidase) is used as a reporter
gene, it is often very difficult to identify the blue cells seen with the
Xgal reaction in a complex tissue. In that case, localizing the Xgal
staining reaction at the EM level (where it can be seen as dark blobs in
the cells) allows one to utilize ultrastructural critera for the cell
identification. There are many other examples.

My own convictions along these lines are clear from the fact that for
the last two years I have offered a graduate course, "Morphology for
Molecular Biologists" here at the University of Michigan. The course is
held once a week (a three hour session), and includes lectures and
demonstrations covering LM, TEM, SEM, a 2-hr "minicourse" in histology,
ultrastructure, and then immunocytochemistry, autoradiography, in situ
hybridization, localization of reporter gene expression, and finally
confocal LM. Grad students from various departments and schools at the
U-M have seemed highly motivated in the course.

So I am betting that the electron microscope does indeed still have
a future.

A. Kent Christensen
Department of Anatomy and Cell Biology
University of Michigan Medical School
akc-at-umich.edu






From: rutledge phil :      prutle1-at-umbc.edu
Date: Thu, 16 Jun 1994 14:31:26 -0400 (EDT)
Subject: Re: technical future in EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopists:
I operate a core facility here at UMBC and we are continually getting
new contracts for EM services. I even get requests from universities
that have EM facilities on campus. EM did take a nose dive but I think
EM is on the way back up. I know this may cause an argument, but even
molecular biologists are starting to see that instead of looking at gels
all the time for information, additional and important information can be
found by doing EM. Yes, I do have a bias against molecular biologists.
When you have been constantly told by molecular biologists that there is
no future for EM, you sort of get biased. I guess they forget that
molecular cell biology is just another TECHNIQUE as is EM.
Have a cold one for EM!
Phil




From: hecub-at-ttacs.ttu.edu (Charles J. Butterick)
Date: Thu, 16 Jun 1994 14:32:16 -0600
Subject: RE: technical future of EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {Chameleon.940616154622.tonygr-at-emlab.mit.edu}

The future of EM also concerns me. Molecular biology is in it's
prime and the investigators seem to be following the trail of grant money
that follows the latest technological fad. Who can blame them?
However, I am still optimistic about the future of all microscopy.
Similar to other labs, we have expanded our LM capabilities and entered
into confocal microscopy. Further, we are also responsible for flow
cytometry. Contracts with industry have been very important to the lab's
fiscal health. Diversification enables us to select the most relevant
technology for the client's needs. Correlation of molecular work to the
morphology will bring the investigators back--in a different way. Immuno,
fluorescence, digital imaging/analysis etc., are all going to be a part of
it. Look at the environmental SEM, or SFM, or AFM. Actually, there is a
lot to be excited about.
What do the material microscopist's think? Eh, Nestor?

************************************************* _______________________
* * | |
* * | _____ ______ |
* Charles J. Butterick (Chuck) * |__| | | |__|
* Electron Microscopy Center * | |
* Department of Cell Biology and Anatomy * ______| |______
* Texas Tech University Health Sciences Center * | __ __ |
* Lubbock, Texas 79430 * |__| | | |__|
* USA * | |
* Phone 806 743-2706 voice * | |
* 806 743-2707 fax * | |
* Email hecub-at-ttacs.ttu.edu * | |
* * __| |__
* * |_______|
*************************************************






From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Thu, 16 Jun 1994 15:59:56 -0500 (CDT)
Subject: TRIM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



A copy of Trim 92 is in the EMC/MSA library. It will be
available at the New Orleans meeting if you want a copy.

Nestor




From: COOK-at-anlemc.msd.anl.gov (Russell E. Cook)
Date: Thu, 16 Jun 1994 15:59:56 -0500 (CDT)
Subject: TRIM code

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: microscopy-at-anlemc.msd.anl.gov

You can contact Peter Baldo at Argonne National Laboratory about TRIM.
Email: PBALDO-at-ANL.GOV
Tele: (708)252-5145 or -7504

Russell E. Cook
Electron Microscopy Center for Materials Research
Materials Science Division
Argonne National Laboratory
Argonne, Illinois 60439
USA






From: bright-at-ENH.NIST.GOV
Date: Thu, 16 Jun 1994 10:46:31 -0400 (EDT)
Subject: TIFF Image files w comments?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We would like to record some information with our images, perhaps in a
header. It is getting to cumbersome to keep matched doc files with their
images. We would like to be able to move and store one file per image and
have that file also contain all of the documentation, instrument
parameters, etc. We would like that information to be in ASCII so that it
can be read easily.

I'm looking into using a TIFF tag such as imagedescription, to designate a
long string or narrative that would describe the image. That string might
look something like this: ((operator Dave Bright)(Instrument JEOL
8500)(Magnification 1000x 5=B5 field width)(date 16 June 1994)(Image_mode
transmission bright field)(comment Sample #3 lot 445x2. 100nm gold coated
after fine polishing. etc. etc.))

I presume we would have to write a small application that would
append/edit/extract this part of the TIFF file. Other tiff file readers
such as NIH Image would ignore it - the actual text would probably be put
after the image, although this should not make any difference.

Anyway, for you image documenters out there: How do you do it? Are there
already utilities around for keeping information along with the images,
other than database type utilities?

Thanks :8o)
Dave

-------------------------------------------------------------
David S. Bright bright-at-enh.nist.gov
Microanalysis Research Group
Bldg. 222 (Chem.) A113
National Institute of Standards & Technology (NIST, formerly NBS)
Gaithersburg, MD 20899-0001 / USA
301-975-3911 (voice), 301-216-1134 (fax)
"A false balance is an abomination to the Lord but accurate weights are his
delight.", Proverbs 11:1






From: bart-at-netcom.com (Harry Bartholomew)
Date: Thu, 16 Jun 1994 21:22:46 -0700 (PDT)
Subject: Re: TRIM code

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


}
} This is hardly microscopy but rather a long shot. I have a colleague here
} looking for a copy of the TRIM (transport of ions in solids) program, which we
} believe to be public domain or otherwise available, for computing
} ion distributions inside a solid after ion implantation. Can anyone help
} direct us in the right direction? We would prefer a Mac or PC version,
} but we also have other workstations so could probably run almost any
} version.
}
} Thanks.
}
} Tony Garratt-Reed
}
}
}

The author is James F. Ziegler
IBM-Research, 28-0
Yorktown, NY 10598


Phone 914 945 2165
Internet: ziegler-at-watson.ibm.com

$5 for a disk and shipping have been suggested.

"TRIM may be copied and distributed freely for academic purpose."





From: hecub-at-ttacs.ttu.edu (Charles J. Butterick)
Date: Fri, 17 Jun 1994 08:19:29 -0600
Subject: Insect prep

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
A client wants to embed ants in plastic for both light and TEM and
has had problems with the exoskeleton embedding adequately in paraffin. We
have no experience with insects. Are there procedures for softening the
chitin? Any help would be appreciated.

************************************************* _______________________
* * | |
* * | _____ ______ |
* Charles J. Butterick (Chuck) * |__| | | |__|
* Electron Microscopy Center * | |
* Department of Cell Biology and Anatomy * ______| |______
* Texas Tech University Health Sciences Center * | __ __ |
* Lubbock, Texas 79430 * |__| | | |__|
* USA * | |
* Phone 806 743-2706 voice * | |
* 806 743-2707 fax * | |
* Email hecub-at-ttacs.ttu.edu * | |
* * __| |__
* * |_______|
*************************************************






From: _ _ _ _ _ MARK W. LUND _ _ _ _ _ :      LUNDM-at-XRAY.BYU.EDU
Date: Fri, 17 Jun 1994 08:47 MDT
Subject: USED MICROSCOPES

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Benci asked about suppliers of used microscopes. Since used equipment
is my favorite kind I have collected a list of companies that include
electron microscopes in thier inventory. Of course this list is not
complete, and I would love to hear about any others:

Cusaco Phone 908-502-9246.

Scientific Equipment and instrument Company 408-428-0464

OOPS! Cusaco's phone is 908-502-0246!! ignore the one above!

The Source 1-800-722-7719

Techlink 408-922-0888

International Equipment Trading Company 708-913-0777

Microscopy Labs 908 747 6228

Conneaut Lake Scientific 814-382-1604

Bid Service 908-775-8300

Regards

Mark W. Lund
Director
MOXTEK, Inc.
Orem UT




From: Mike Nesson :      nessonm-at-BCC.ORST.EDU
Date: Fri, 17 Jun 1994 15:48:26 -0700 (PDT)
Subject: Re: Insect prep

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {Chameleon.940617133055.tonygr-at-emlab.mit.edu}


On Fri, 17 Jun 1994, Charles J. Butterick wrote:

} Greetings,
} A client wants to embed ants in plastic for both light and TEM and
} have no experience with insects. Are there procedures for softening the
} chitin? Any help would be appreciated.

You might look at a recently published method in which the cuticular or
waxy surface is primed by treatment with gamma-glycidoxypropyl
trimethoxysilane prior to resin embedment in LR White. the paper was in
Microscopy Research and Technique 21:355-360 (1992). The silane is
available from Polysciences,Inc. (800)-523-2575.

Good luck,
Mike Nesson






From: ars-at-mcs.com (Allen R. Sampson)
Date: Fri, 17 Jun 1994 17:59:21 -0500
Subject: Re: Used SEM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0qEmzI-000BcgC-at-mercury.mcs.com}
X-Sender: ars-at-mcs.com
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

In article {9406161718.AA21030-at-inet-gw-1.pa.dec.com}
jbenci#a#hub#d#eng#d#wayne#d#edu.MSE-at-mse.engin.umich.edu ("John Benci") writes:
} From: jbenci#a#hub#d#eng#d#wayne#d#edu.MSE-at-mse.engin.umich.edu ("John Benci")
} Subject: Used SEM's
} Date: 16 Jun 1994 12:29:35 U




From: ars-at-mcs.com (Allen R. Sampson)
Date: Fri, 17 Jun 1994 17:59:21 -0500
Subject: Re: Used SEM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Thanks for your help.

} John Benci
} Materials Science & Engineering
} Wayne State University

} e-mail address: jbenci-at-eng.wayne.edu


} ------------------ RFC822 Header Follows ------------------
} Received: by mse.engin.umich.edu with SMTP;16 Jun 1994 12:29:09 U
} Date: Thu, 16 Jun 94 12:15:18 EDT
} From: jbenci-at-hub.eng.wayne.edu (John Benci)
} Message-Id: {9406161615.AA29719-at-ss0.NIS.ss0}
} To: microscopy-at-anlemc.msd.anl.gov
} Subject: Used SEM's


A gentleman by the name of Clark Houghton, located in Ohio, would be the best
and closest source I know of. Like me, he offers maintenance services on a
wide range of electron microscopes. Unlike me, he also deals in used,
refurbished
instruments and also has leasing options. There are others offering used
instruments, but if I were looking for a used instrument, he's probably the
only one I would talk to.

Clark Houghton
Phone (513) 927-5373
FAX (513) 927-5557

BTW, although I don't generally deal in used equipment, I often help my
customers find new homes for their old stuff. None that I know of currently
have equipment available, but I expect at least a couple of machines on the
market in the next 6 months to a year. If your needs are not immediate, send
me email with a more specific listing of your requirements.

Just to clarify - I have no financial relationship to Mr. Houghton, and only a
passing friendship. He is well known and respected in the field.


Allen R. Sampson

Advanced Research Systems
317 North 4th. Street
St. Charles, IL 60174

PH (708) 513-7093 FAX (708) 513-7092
Internet: ars-at-mcs.com Compuserve: 71271,330

repair and maintenance services for analytical instrumentation





From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: 17 Jun 1994 18:31:44 -0600
Subject: Re: EM - Facility size and # of users?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Could you please tell me if you are a state or private institution. Who bought
the microscopes, who pays the tech salaries and who buys the supplies at
priori. Finally who pays for the service contrast of each instrument. Do your
charges to customers reflect these categories? How much let us say you would
charge for embedding a piece of heart cuttting at three levels and making 30
pictures (5X7 inches) 10 at low mag (1-3) 10 at 5-20 and ten at 20-100?
Thanks.

***** ************ ************** ************
*Cesar D. Fermin, Ph.D |Fax (504) 587-7389
*Tulane Medical School |Answ. Mach.(504) 584-2618
*Pathology/SL79 |Secretary (504) 584-2436
*New Orleans, La 70112 | Lab (504) 584 2521
***** ***************** ***********************
_______________________________________________________________________________
Cc: Microscopy Group

In answer to Dwight Beebe's additional questions:

1. Our laboratory servers approximately 80 major users (repeat more than
2X a year). 9 from our department, 41 from other departments within our
institution, 24 from outside our institution (4 from industry), and 6
from foreign countries.

2. We charge less for departmental users because our department pays part
of the service contract on the instruments.

3. We do not charge less based on user expertise. We provide initial
training in any aspect of microscopy the user desires. After that we
charge a fee for technical time if one of our staff is involved in the
microscopy.

4. As outlined in a previous communique (Nestor- do we have a FAQ
section?), we breakdown our fees for each individual technique done and
cKcharge what it costs us to complete the task.Determining these fees is
a major headache, but simplifies administration later and avoids hurt
feelings since all are charged based only on what they use. We of course
do not nit pick (i.e. charge for each grid, each cc of osmium, etc).
Rather we have an average cost for negative stain, cost for
cryosectioning, cost for cyroEM, etc.

5. Microscope usage is on a sign up reservation basis. Inexperienced
users can only sign up during hours when resource staff are available for
consultation. Once we (as a group) are convinced a user can handle
themselves we make the microscopes available to them 24 hours a day.
Rarely, but it has happened, we encounter someone who is either incapable
or unwilling to use the laboratory in a responsible way. These people are
denied all access.

6. Marcelle Gillot's summary of the problems of determining users fees is
excellent and is almost identical to the constraints we use for
determining fees. However, our legal department has a slightly different
interpretation of allowable charges for outside, non-grant supported
work. They advise that it is allowable to undercut the cost of locally
available commercial services as long as you can document (and that may
be a problem) that you do so while still recouping all costs of the work
without relying on subsidies (goverment grants, institutional support for
service contracts, salaries, building maintenance and rent etc.) to cut
your cost. In other words, if
you are willing to take less profit margin or pay your labor sweat shop
wages you can charge less than your commercial competition. You just need
to prove that you and your competition are playing on a level playing
field.

I hope this is helpful.
I would support a movement within MSA to survey and document the issues
involved with fee for service operation. Business is clearly something
most of us where not traine in and really don't want to spend our time
on- however the world being what it is today..........
Regards
Jay Jerome





From: diandra-at-unlinfo.unl.edu (diandra lesliepelecky)
Date: Sun, 19 Jun 1994 11:04:09 -0500 (CDT)
Subject: Skeletal Carbon TEM grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a procedure for producing skeletal carbon TEM grids
with holes sufficiently small to capture particles ranging in size
from 200 nm to 30 nm?

Thanks.

--
Diandra L. Leslie-Pelecky
Center for Materials Research and Analysis PH: (402) 472-9178
University of Nebraska FAX: (402) 472-2879
Lincoln, NE 68588-0113 diandra-at-unlinfo.unl.edu




From: Robert Fisher :      rmfisher-at-u.washington.edu
Date: Sun, 19 Jun 1994 10:48:48 -0700 (PDT)
Subject: EDS Cryogenics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I wish to confer with someone knowledgeable about
the cryogenic aspects of EDS detectors.
e.g., loss of insulating capacity of the Dewar,
effects of warmup - quantitative, i.e., resolution with
time and temperature- activation energy of the kinetics,
effects of operating at dry ice -acetone temperatures.

A really good reference will suffice.

Please reply direct. I will prepare digest of any information
that I receive for the list.

R.M. Fisher
rmfisher-at-U.Washington.edu




From: {mwinton-at-nosc.mil}:ddn:wpafb
Date: 6-16-94 5:35pm
Subject: Software / Font needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9406201136.AA17097-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: Software / Font needed
Orig-Author: {mwinton-at-nosc.mil (Michael J. Winton)}:ddn:wpafb
-----------------------------------------------------------
I apologize in advance for a post which is somewhat related to
microscopy but not very closely. Please skip this message if
that bothers you.





I am looking for a font for the Macintosh which has overlined
numbers (Miller indices) built in. I have been searching and
searching, and am tired of "drawing" the indices and pasting them
into my documents. I'm sure that someone out here must have
such a font. A public domain font editor would also be useful--
then I could create such a font and make it available to others.
I have yet to find a font editor in an FTP site--font utilities are
usually useful for little more than cataloging. I thought that
there would be enough scientists in this newsgroup who have to
deal with this regularly that one of you might know of something.

Please let me know if you can help.



Michael Winton (mwinton-at-nosc.mil / -at-zazen.lbl.gov)










From: yang-at-snmail.jsc.nasa.gov
Date: 6/19/94 11:07 AM
Subject: Skeletal Carbon TEM grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Does anyone have a procedure for producing skeletal carbon TEM grids
with holes sufficiently small to capture particles ranging in size
from 200 nm to 30 nm?

Thanks.

--
Diandra L. Leslie-Pelecky
Center for Materials Research and Analysis PH: (402) 472-9178
University of Nebraska FAX: (402) 472-2879
Lincoln, NE 68588-0113 diandra-at-unlinfo.unl.edu





From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Tue, 21 Jun 1994 16:48:22 +1200
Subject: LM Fibrocartilage stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear Light Microscopists,
Does anyone know of a specific LM stain for fibrocartilage? A student is
doing a study on a part of the skull and joints and would like some help.
Any suggestions?

Thank you
Piotr Swierczynski

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Tue, 21 Jun 1994 17:07:32 +1200
Subject: Biorad Sputter-coater targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear EM people,
Does anyone know of any company (other Fisons UK) manufacturing gold or
gold/palladium ring targets for Biorad sputter-coaters? We have just been
informed (after waiting two weeks and having being informed that it was on
its way) that we have a 3-4 week wait for a new one from Fisons UK, a
slight problem as due to an "administrative oversight" there is no spare.
Any suggestions for an alternative source in future would be gratefully
received.
Yours faithfully
RE

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: timonf-at-earth.ruu.nl (Timon Fliervoet)
Date: Tue, 21 Jun 1994 09:13:23 +0200
Subject: Re: Software / Font needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9406210713.AA17730-at-escher.earth.ruu.nl}
X-Sender: timonf-at-escher.earth.ruu.nl
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

In Word for Macintoshes, Miller indices are easy to make using the formula
editor.

For those using Macs and don't know how to use the formula editor I can
send a small Word 5.1 document with some examples of Miller and Weber
indices. (BTW Word for Macs is much easier to use than WordPerfect!!).

Cheers, Timon

------------------------------------------------
Timon Fliervoet, Timonf-at-earth.ruu.nl, Faculty of Earth Sciences, Department
of Geology, Utrecht University, P.O. Box 80.021 3508 TA Utrecht, the
Netherlands. tel: (0)30 - 535054, fax: (0)30 - 537725






From: Eliesh O'Neil :      eliesh.oneil-at-gtri.gatech.edu
Date: Sun, 19 Jun 1994 08:49:05 -0500 (EST)
Subject: RE: Biorad Sputter-coater targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Richard,

At least one supplier here in the U.S. -

SPI Supplies carries the targets you are looking for and they should fit the
Bio-Rad instrument. They are located in West Chester, Pennsylvania, USA.
Tel # 1-800-242-4774 and FAX # 1-215-436-5755. Good luck!


In message Tue, 21 Jun 1994 17:07:32 +1200,
richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood) writes:

} Dear EM people,
} Does anyone know of any company (other Fisons UK) manufacturing gold or
} gold/palladium ring targets for Biorad sputter-coaters? We have just been
} informed (after waiting two weeks and having being informed that it was on
} its way) that we have a 3-4 week wait for a new one from Fisons UK, a
} slight problem as due to an "administrative oversight" there is no spare.
} Any suggestions for an alternative source in future would be gratefully
} received.
} Yours faithfully
} RE
}
} Richard Easingwood
} Department of Anatomy and Structural Biology,
} P.O. Box 913
} University of Otago,
} Dunedin, New Zealand.
} Fax:64-3-479 7254
} Telephone:64-3-479 7301
}




From: David Henriks :      73531.1344-at-CompuServe.COM
Date: 21 Jun 94 11:08:22 EDT
Subject: Biorad Sputter-coater targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A very reputable company in the UK that handles a wide range of sputtering
targets is:

Testbourne Ltd.
Unit 12, Hassocks Wood
Stroudley Road
Basingstoke, Hampshire RG24 ONE
England

TEL: (256) 467-055
FAX: (256) 842-929

Contact: Ted Mihill

I understand that they can ship stock targets in 24 hours and custom targets in
2-4 weeks. Could be worth a try. Good luck!

Best regards-

David Henriks
South Bay Technology, Inc.
TEL: 714-492-2600
FAX: 714-492-1499





From: jester-at-crnjjsgi.swmed.edu (James V. Jester)
Date: Tue, 21 Jun 1994 11:03:56 -0500
Subject: Journal of Anatomy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have the address, phone number, FAX, Email for the
Editorial Office of Journal of Anatomy.

_____________________________________________________________________
| | |
| James V. Jester | Dept Ophthalmology |
| jester-at-crnjjsgi.swmed.edu | UT Southwestern Medical Center |
| TEL (214)648-7215 | 5323 Harry Hines Blvd |
| FAX (214)648-2382 | Dallas, TX 75235-9057 |
|__________________________________|__________________________________|





From: eclausnz-at-eye1.eye.ufl.edu
Date: 21 Jun 1994 11:57:33 EST
Subject: LM STAIN FOR MITOTIC ACTIVITY

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {MAILQUEUE-101.940621115228.256-at-eye1.eye.ufl.edu}
To: microscopy-at-anlemc.msd.anl.gov

Several of our people are working with whole mounts of retinas removed from
cow eyes, and are looking for a stain that will differentiate between
dividing cells and inactive ones. We could use any suggestions for LM or
fluorescent stains that do not require embedding and sectioning. Thanks!
Evelyn Clausnitzer
U of Florida
Ophthalmolgy




From: :      MELLIOTT-at-prl.pulmonary.ubc.ca
Date: Tue, 21 Jun 1994 11:00:03 +0800PST
Subject: LM fibrocartilage stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Piotr Swierczynski was asking for specific stain for fibrocartilage.
Don't think there is one-no-one here knows of one. Suggest you stain
for collagen-thick bands in amongst cartilage would differentiate it.
Possible stains would be Wright's, Masson-Trichrome or H and
E/alcian blue double stain. Could also try an alizarin red S and
Toluidine Blue O stain (Dawson 1926, Stain Techn. 1:123-4; Williams
1941, Stain Techn. 16:23-5).
Let us know what you find out-which one works better.

Mark Elliott,
UBC-Pulmonary Research Lab,
St.Paul's Hospital
Vancouver BC




From: :      MELLIOTT-at-prl.pulmonary.ubc.ca
Date: Tue, 21 Jun 1994 11:04:44 +0800PST
Subject: stain for mitotic activity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Evelyn Clausnitzer was asking about stains for whole mounts for
mitotic activity-try DAPI, Hoechst or acridine orange. They all work on
tissue cultured cells which have not been embedded. Should be able
to tell mitotic figures easily. Check with Molecular Probes to see if
there are others, but these are fairly easy to use and look great.

Mark Elliott
UBC-Pulmonary Research Lab
St. Paul's Hospital
Vancouver BC Canada




From: rutledge phil :      prutle1-at-umbc.edu
Date: Tue, 21 Jun 1994 16:15:14 -0400 (EDT)
Subject: LM stain for fibrocartilage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Piotr Swierczynski:

Have you tried just a basic H&E stain? It works well
on fibrocartilage from a human intervertebral disk. Is the brain damaged?
If so, have you tried: "Foot's Ammoniated Silver Carbonate Method"? From
Foot (1924). It will show: vascular reticulum, tumor cells , and
connective tissue around a brain tumor. Here is the procedure if you want
to try it:

Fix: formalin (1:10) recommended; Cajal's FAB and Bouin's fluids also
satisfactory.

Embed: in paraffin

Solutions:
A. Ammonical silver carbonate: To 10 ml of 10.2% AgNO3 add conc.
NH4OH, drop by drop until precipitate is almost redissolved; then
add 10 ml of 3.1% Na2CO3, concentrated formalin, 1 ml;
water, 100 ml.
B. Reducing agent: 1% Na2CO3, 3 ml; conc. formalin, 1 ml; water,
100 ml.
C. Intensifying solution: oxalic acid, 2 gm; conc. formalin, 1 ml;
water, 100 ml.

Staining schedule:
1. remove paraffin from mounted sections and treat for 24 hr at room
temp. in a mixture of pyridine and glycerol (2:1 by volume).
2. rinse in 95% etoh, then in water.
3. impregnate for 2.5 hr at 40 C in silver sol. A.
4. wash in distilled water.
5. reduce 5 min in sol. B.
6. wash in tap water.
7. tone 5 min in 0.2% gold chloride.
8. wash in tap water.
9. intensify by treating 5 min in sol. C.
10. rinse in tap water.
11. fix in 5-10% Na2S2O3.5H2O.
12. wash in tap water.
13. dehydrate.
14. cover in balsam or synthetic resin.

Note:
The stain is not recommended for normal brain. It has considerable
value in demonstrating connective tissue reaction around tumors
if reticulin fibers are laid down by this connective tissue. If
FAB fix is used, the reticulin staining is said to be suppresed.

Good luck!
Phil




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Tue, 21 Jun 1994 22:01:48 -0500 (CDT)
Subject: Announcing Tribology Listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


June 19, 1994

ANNOUNCEMENT: Tribology and Coatings Technology Listserver/Mailreflector

Dear Colleagues:

With this EMail message we announce the startup of a Tribology
Listserver/Mailreflector System. The purpose of this system
is to allow the scientific community a centralized Internet
address to which questions/comments/answers in the various
fields of Tribology and Coatings Technology
can be rapidly distributed to a list of interested individuals.

For the purposes of this discussion forum, Tribology should
be considered to include all aspects of the study of coatings
which involve friction, wear, lubrication, corrosion, strength,
stability, design, and fabrication of materials and/or processes which
involve the interaction of a surface (or surfaces) with its environs.

Thus, the purpose of this announcement of the
tribology listserver is to provide an initial impetus
and forum for discussion of problems which can benefit the
presentation/discussion by the tribology and coatings community as a whole.

There are no charges to become a member of the subscription
list nor are there any charges for usage, except for the
request that you actively participate in any discussion to
which you have a question, comment and/or contribution.
All individuals regardless of their affliation (commerical,
educational, or government) are welcome to join.

To register with the system

1.) Send an Email message to "LISTSERVER-at-ANLEMC.MSD.ANL.GOV"
within the body of the Email message include the following
line:

"Subscribe TRIBOLOGY UserName-at-EMailaddress"

where "EMailaddress" is the electronic mail address of your
host computer, and "UserName" is the username by which you are
registered on that system.

PLEASE NOTE: The word TRIBOLOGY is important in
your subscription request as this LISTSERVER software
processes Email for several different mailing lists.

2.) Within a day or so you will receive a confirmation test
message which tests the address that you have supplied
to the ListServer. Upon receipt of a reply from the
subscriber (i.e. UserName-at-EMailaddress) to this test message
your name will be added to the mailing list. You will then
automatically receive copies of "ALL" Email sent to this
system.

3.) Anyone may post messages to this list, however, only
subscribers will receive copies. One may post messages/comments
to this list using any conventional Email system by sending a
message to:

"TRIBOLOGY-at-ANLEMC.MSD.ANL.GOV"

4.) To remove your UserName from the list send an Email message to
LISTSERVER-at-ANLEMC.MSD.ANL.GOV containing the line:

"Unsubscribe TRIBOLOGY UserName-at-EMailaddress"


=======================================================

This Tribology Listserver suplements the ANL
Microscopy ListServer and MSA (Microscopy Society of America)
electronic bulletin board system, and the EMMPDL (Microscopy and Microanalysis
Public Domain Library) as a means of electronic communications which are
also accessible over Internet as well as conventional
telecommunications (i.e. modem) lines. Details of these are available
upon request from EMMPDL-at-ANLEMC.MSD.ANL.GOV.

Please feel free to distribute this announcement to any individuals
or groups whom you think may be interested in participating.

Also, as this is an experimental and FREE, service to the commmunity
please understand that some system problems will occur. They will
be dealt with as (my) time permits but it is being done on a purely
voluntary (& masochistic) basis.

==========================================================================
Nestor J. Zaluzec
Materials Science Division , Bldg - 212
Argonne National Laboratory, Argonne, Ill. 60439 USA

Tel: 708-252-5075 Fax: 708-252-4798
Email: Zaluzec-at-anlemc.msd.anl.gov
===========================================================================




From: {temcom-at-engin.umich.edu}:ddn:wpafb
Date: 6-21-94 1:29am
Subject: Re: Word vs Wordperfect via Software / Font needed discussion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9406221358.AA27889-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb
Subj: Re: Software / Font needed
Orig-Author: {Brian Gregory Demczyk {temcom-at-engin.umich.edu} }:ddn:wpafb
-----------------------------------------------------------
No one in their right mind would use Word Perfect instead of Microsoft
Word! Of course one is able to overstrike in the latter!













From: Donald L. Grimes :      74250.331-at-CompuServe.COM
Date: 22 Jun 94 10:22:38 EDT
Subject: Polaraid Contest

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Group -
Per a recent request, the winners of the Polaraid International
Instant Photomicrography Competition are:
Grand Prize: Dr. Gerald T. Baker, Mississippi State Univ.
B&/W Light Micrography: Floyd E. Alberts, Ford Motor Company
Microscopy Techniques: John Georgiou, University of Toronto
Material Sciences: Vito Giannini, Italcementi SpA (Italy)
Student: Jim Wetzel: Clemson Univ.
Color Light Micrography: Wutian Wu, Eastern Virginia Medical School
Should any wish a copy of the full 3-page announcement, send me
your fax number and I will be pleased to supply.
I expect that we will do an article on the full winners list in a
future issue of Microscopy Today.
Regards,
Don Grimes, Microscopy Today






From: yang-at-snmail.jsc.nasa.gov
Date: 6/19/94 11:07 AM
Subject: Skeletal Carbon TEM grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Does anyone have a procedure for producing skeletal carbon TEM grids
with holes sufficiently small to capture particles ranging in size
from 200 nm to 30 nm?

Thanks.

--
Diandra L. Leslie-Pelecky
Center for Materials Research and Analysis PH: (402) 472-9178
University of Nebraska FAX: (402) 472-2879
Lincoln, NE 68588-0113 diandra-at-unlinfo.unl.edu






From: Carlo Montemagno :      cmontema-at-hawk.ce.nd.edu
Date: Tue, 21 Jun 1994 12:33:06 -0700
Subject: microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I need to purchase an inverted epi microscope to conduct membrane kinetic
studies. Because of the experimental setup I can't use immersion optics. I have
about 22K to spend and I already have a good quality video camera. Are there
any thoughts on which product lines are the best and recomentations for
objectives?

Thanks

Carlo Montemagno





From: vierreth-at-rorc.usbm.gov
Date: Wed, 22 Jun 1994 11:49:41 -0500 (CDT)
Subject: Re: microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Tue, 21 Jun 1994, Carlo Montemagno wrote:

} I need to purchase an inverted epi microscope to conduct membrane kinetic
} studies. Because of the experimental setup I can't use immersion optics. I have
} about 22K to spend and I already have a good quality video camera. Are there
} any thoughts on which product lines are the best and recomentations for
} objectives?
}
} Thanks
}
} Carlo Montemagno
}
}
Try a Nikon, if you want a rotating stage a Nikon will not do.
Clarissa







From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 22 Jun 1994 14:51:44 U
Subject: Re-Overbars in MS Word

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subject: Time:2:46 PM
OFFICE MEMO Re:Overbars in MS Word Date:6/22/94
It is possible to print nearly any selected character centered
over nearly any other character by using, in combination, the
'Overstrike' and 'Superscript' typesetting commands provided in
Microsoft Word. These commands are described in the
section on producing formulas (which appears on pages 98 to
105 in the instruction manual for MSW V4.0).
As an example, working with New York font, the following
command string will produce a capital 'X' with a bar neatly
centered above it:
.\O.\AC(X,.\S.\UP11(_)),
while the command string:
(.\O.\AC(1,.\S.\UP11(_))0.\O.\AC(1,.\S.\UP11(_)))
produces the set of Miller "bar one-zero-bar one".
In these strings, the special character (.\) telling MSW to
enter the typesetting mode is produced by holding down the
command and option keys while typing a backslash (\).
Command strings must be prepared by activating the
"Show-#166#" mode , and the final printed format can be
displayed by the "Hide-#166#" mode (obtained in the Edit menu
or toggled by pressing the command and 'Y' keys
simultaneously).
Now, you may complain that it is inconvenient to type long
command strings such as these as frequently as might be
needed in many manuscripts. However, the process can be
simplified by first typing the manuscript using a dummy
variable (e.g. VX for vector X, 1B for overbar one, etc.). When
completed in this manner, prepare the command string for a
particular symbol and save it on the clipboard (command-c).
Then use the Find utility (command-F) to find the first
occurrence of the corresponding dummy variable, click in
the ULH square to release the Find window, replace the
dummy variable with the command string (command-V),
find the next occurrence of the dummy variable
(option-command-A), replace it, and so proceed through
the entire manuscript.
Try it, you'll find it works rather well. You will also find
the full set of typesetting commands to be very useful
for inserting formulas into manuscripts - they require a little
practice, but turn out to be quite convenient and versatile,
once you get familiar with the system.






From: {dlc-at-owlnet.rice.edu}:ddn:wpafb
Date: 6-22-94 12:14pm
Subject: overstriking...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9406221911.AA29215-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb
Subj: overstriking...
Orig-Author: {"Daniel L. Callahan" {dlc-at-owlnet.rice.edu} }:ddn:wpafb
-----------------------------------------------------------

If you find out how to overstrike in Word (Mac), please post to the
microscopy listserver so all will know! I too have hunted this down
unsuccessfully.

Daniel L. Callahan
Department of Mechanical Engg. and Materials Science
Rice University
dlc-at-owlnet.rice.edu






From: Ian Hall :      hall-at-me.udel.edu
Date: Wed, 22 Jun 1994 17:06:58 -0400 (EDT)
Subject: Re: Word vs Wordperfect via Software / Font needed discussion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Scott,
To overstrike in Word for the Mac is staggeringly simple. To put
e.g., a slash through a zero similar to one of the Scandinavian
characters, you should i) hold down COMMAND and OPTION together and type a
backslash ii) type an o (for overstrike strangely enough) iii)) type a LH
bracket iv) type the two characters, separated by a comma v) close the
brackets. (Of course the Scandinavian letters are available anyway so you
would not bother doing this particular example)
Congratulations, you have just overstruck! If you want to put a
bar above an index, just make the second character a superscript hyphen.

Happy overstriking
Rick Hall
Mat.Sci., U. of Delaware






From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Wed, 22 Jun 1994 16:43:11 -0500 (CDT)
Subject: It's cool off time gang

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Subscribers.....

Enough on Word vx. Word Perfect and let's get back
to Microscopy.....

Nestor
Your friendly neighborhood SysOp




From: Dwight Beebe :      beebed-at-ERE.UMontreal.CA
Date: Wed, 22 Jun 1994 19:47:39 -0400 (EDT)
Subject: Re: Word vs. Wordperfect

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Okay folks,
I'm more than a little suprised that Nestor hasn't chimed in here
and I might be out of place, but this is _NOT_ the place to either
discuss the pros and cons of a particular piece of word processing
software, particularly when it has no direct bearing on microscopy, nor
is this forum the place to flame someone for the signal to noise ratio of
their posts. I have been the unfortunate participant (vicarious) in
other flame wars on different forums and they serve _NO_ purpose. If you
find the comments of someone offensive and wish to respond, please do so
in private. Please, please, don't let this highly useful and informative
discussion group dissolve into acrimonious shoot-outs over message
content or perceived attitude.
Thanks, I'll now fade back to my usual semi-transparent level of
participation.

Dwight U. Beebe
Institut de recherche en biologie vegetale beebed-at-ere.umontreal.ca
Universite de Montreal Voice:514-872-4563
4101, rue Sherbrooke est FAX:514-872-9406
Montreal, PQ H1X 2B2 Canada







From: Dwight Beebe :      beebed-at-ERE.UMontreal.CA
Date: Wed, 22 Jun 1994 19:52:05 -0400 (EDT)
Subject: Re: It's cool off time gang

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Aaaarrrgghh!
Guess that'll teach me to read all my mail before inserting my
long nose into the fray. Sorry for the wasted bandwidth.

Dwight






From: timonf-at-earth.ruu.nl (Timon Fliervoet)
Date: Thu, 23 Jun 1994 08:38:58 +0200
Subject: Re: overstriking...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9406230638.AA23830-at-escher.earth.ruu.nl}
X-Sender: timonf-at-escher.earth.ruu.nl
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

To all who are more interested in word typesetting:

to put a bar over a 1, the nicest results (I think) are given by:

1 command-option-\D command-option-\ba5()command-option-\S command-option-\
up7(-)

NB the command key is the key marked with an apple sign
NB2 hold down the option, command and backslash (\) key simultaneously
(this will activate the formula editor).
NB3 it might be a bit longer then option-command-\O(1,_) but can be put in
a macro

Cheers Timon

------------------------------------------------
Timon Fliervoet, Timonf-at-earth.ruu.nl, Faculty of Earth Sciences, Department
of Geology, Utrecht University, P.O. Box 80.021 3508 TA Utrecht, the
Netherlands. tel: (0)30 - 535054, fax: (0)30 - 537725






From: peling-at-amnh.org (Peling Melville - Interdepartmental Facilities)
Date: Thu, 23 Jun 1994 08:20:11 -0500
Subject: The word vs. wordperfect debate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi folks,

I just want to thank both Nestor & Dwight for bringing out the fact that
this IS a microscopy list. I agree wholeheartedly that this is NOT the
place for discussions (shall we say) on things such as word vs.
wordperfect. Let's get back to microscopy please.

Just venting.

Thanks,
Peling Melville

--------------------------------------------------------------
Peling Melville peling-at-amnh.org
Interdepartmental Laboratories
American Museum of Natural History






From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 23 Jun 1994 09:47:01 U
Subject: Overstrike FONT

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reply_ Overstrike FONT
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX: (313)936-3352
jfmjfm-at-umich.edu or John_Mansfield-at-mse.engin.umich.edu
A company in Ann Arbor, Allotype Typographics sells two fonts that are great
for Miller indices. They are called Haber and Thompson, they are based on
Helvetica and Times respectively. I have been using these fonts for about 5
years now and have been very happy with them.
Allotype can be reached at 313-480-3666.
I am just a satisfied customer.
Jfm.






From: KINGSLAND, Arlene :      KINGSLAND-at-paprican.ca
Date: Thu, 23 Jun 1994 10:02:31 EST5EDT
Subject: Wood pulp fibres sticking to slides.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.940623100231.480-at-pap386.paprican.ca}

Hello Everyone,

We are trying to use a procedure that was established over 20
years ago. One of our problems is that a chemical was used to treat
the glass slides in order to prevent the wood pulp fibres from
sticking to them during drying. The chemical used was
dichlorodimethylsilane which is considered highly toxic. We would
like to hear any suggestions for an alternative way to prevent the
fibres from sticking.

Thanks,

Arlene Kingsland.




From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Thu, 23 Jun 1994 09:20:13 -0500
Subject: what's appropriate?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
I would like to say that the discussion on overstriking has been
very helpful. While I agree that those parts of the discussion that
"flamed" were not needed, the main discussion had a lot of content, and I
don't see why this disscusion should not be allowed to continue on this
group.

I bet word processing is the ONLY technique that EVERYone on this
list uses. Word processing certainly is an indespensible technique for the
professional microscopist. I guess it is reasonable to have some limits:
certainly flames are unwanted, and so too discussions of an ethically
dubious nature might be shunned, such as how microscopists can avoid income
taxes, and finally discussions that are wholly irrelevant, such as on
making the best cheese fondue could be canned. My view is that word
processing is relevant for microscopists to talk about. I might add, as a
light microsopist in a biology department, there are many subjects covered
on the list that I ignore. Three cheers for clear subject lines.

A veiw from the trenches.

Tobias Baskin

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Tobias I. Baskin Baskin-at-biosci.mbp.missouri.edu
University of Missouri * Division of Biological Sciences * 109 Tucker Hall
Columbia, MO 65211 USA voice: 314-882-0173 fax: 314-882-0123
___ ____ ^ ____ _____
/ \ / / \ / \ /
/ | / / \ / /
/___ / /__ /_____\ / /__
/ / / \ ( /
/ / / \ \ /
/ /____ / \ \____/ /_____






From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Thu, 23 Jun 1994 10:04:12 -0500 (CDT)
Subject: Microscopy Microanalysis & Word Processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Subscribers

To be fair to the individual who started the discussion
thread on "Miller Indices" (not word processors!), the originator of the
message, first sent the text to me and I was asked if it was appropriate
for this forum. Since it was a serious question, and one which was
related to microscopy and he was clearly in need of an answer (& and
a listing of the various options for solving his problem) I concurred
and told him to post the question to this forum. It's now run it's
course and unless there is something else to add specifically on
Miller Indices that has not been covered let's move on.

I will continue to be a behind the scene's filter for anything which
the subscribers to this list may wish an opinion on prior to posting and a
traffic cop when things appear to be getting shall we say "out of hand".
Feel free to abuse my screen all you like. You will
neither be the first and I'm sure you will not be the last, and as you are
all aware the electronic trash can is only a button away.

Nestor





From: meh-at-aretha.jax.org (Margaret Hogan)
Date: Thu, 23 Jun 1994 11:04:56 -0400
Subject: bulletin board content..

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199406231505.LAA24785-at-aretha.jax.org}
X-Sender: meh-at-aretha.jax.org
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I would have to agree with Tobias Baskin's comments on acceptable content of
this bulletin board. As long as the material is something used by the
readers on a professional level, why not include it as fair game? In an
open system, such as this, it's up to the contributors to remain
professional, and keep in mind what may be of general interest to the rest
of the readers.

I find this forum interesting to read, and it's provided a wealth of
information that I've squirreled away for future reference. I'd really like
to see it kept open...

Thanks


Margaret E. Hogan, PhD
Electron Microscopy Service
The Jackson Laboratory
600 Main Street
Bar Harbor, Maine 04609
(207) 288-3371 #1450
(207) 288-5079 FAX
meh-at-aretha.jax.org





From: P.V.Hatton-at-sheffield.ac.uk
Date: 23 Jun 94 18:47:32
Subject: Microscopy, not word-processors!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {MAILQUEUE-101.940623184732.288-at-UNDERBANK.shef.ac.uk}
To: microscopy-at-anlemc.msd.anl.gov

Hi, I would have kept quiet but for some users suggesting that it is
a good idea to exchange trivia about word processing. I joined a
microscopy board, not a software panel - I get a lot of useless mail
and I don't need a pile of messages about word processing. May I
politely suggest that people who want to talk about anything except
microscopy on this valuable service should form their own club.

Votes for and against?

Paul Hatton
University of Sheffield






From: llsutter-at-mtu.edu (Larry Sutter)
Date: Thu, 23 Jun 1994 14:26:33 -0400
Subject: Microscopy, not word-processors!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



As long as everyone is getting in the act, I will also jump in. I
am a Mac user and I subscribe to a variety of newsgroups, many of which
specialize in discussions of topics such as WP programs. I suggest that
other Mac users seek out such discussion groups and PC users seek like
discussions. Let's talk scatterin' events here...


Larry Sutter
Michigan Technological University
Department of Metallurgical & Materials Eng.
1400 Townsend Dr.
Houghton, MI 49931






From: waheeschen-at-dow.com (Bill Heeschen (517)636-4005 PMRC/ASL)
Date: Thu, 23 Jun 1994 14:52:11 -0400
Subject: Photometrics Star 1 CCD camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Private response only!!!

Does anyone have/know of a Photometrics Star 1 thermoelectrically-cooled CCD
camera in good condition which is available for purchase? They are apparently
no longer commercially available (that's why I am bothering this list about it.)


Bill
==========================
Bill Heeschen/Analyt. Sci.- Mat. Char. -----
1897-F Building / The Dow Chemical Co. --- ---
Midland, MI 48667 U.S.A. --- D O W ---
phone: (517)636-4005 fax: (517)636-5453 --- ---
Email: waheeschen-at-dow.com -----
========================== R




From: Paul Webster :      Paul_Webster-at-quickmail.yale.edu
Date: 23 Jun 1994 16:41:41 -0400
Subject: Is EM art?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subject: Time: 4:42 PM
OFFICE MEMO Is EM art? Date: 6/23/94
Are we scientists are artists? If the answer is the first then why are there
so many micrograph competitions?






From: rutledge phil :      prutle1-at-umbc.edu
Date: Thu, 23 Jun 1994 17:20:14 -0400 (EDT)
Subject: Re: Is EM art?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On 23 Jun 1994, Paul Webster wrote:

} Subject: Time: 4:42 PM
} OFFICE MEMO Is EM art? Date: 6/23/94
} Are we scientists are artists? If the answer is the first then why are there
} so many micrograph competitions?

} To answer this.
EM is an art! You either have it or you don't!

Phil




From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Thu, 23 Jun 1994 17:10:44 -0400 (EDT)
Subject: Re: Is EM art?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

EM Bulletin Board {microscopy-at-anlemc.msd.anl.gov}
In-Reply-To: {199406232055.AA11541-at-alsys1.aecom.yu.edu}
Message-Id: {Pine.3.07.9406231743.A6730-c100000-at-alsys1}
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; charset=US-ASCII

} OFFICE MEMO Is EM art? Date: 6/23/94
} Are we scientists are artists? If the answer is the first then why are there
} so many micrograph competitions?
This question has no simple answer. For starters, the question of who
(or what) an artist is may need to be answered. Answering the question,
"What is a scientist?" is not easy, but at least science is a
collection of fairly well pre/proscribed paradigms. Art, on the other
hand, is a free-for-all.
It is now generally accepted that photographs may be art; I say generally
accepted because there are some very conservative/reactionary/retrograde
critics who refuse photography art status, but if we relegate these people
to the status of bonkers, then we'd all agree (as any reasonable person
would) that photographs can be art.
So what makes a micrograph, a specific class of photograph (although this
definition must be broader to include video, computer representations, etc.) a
work of art? Let's look at a more general
question, what makes (an object) art? Or a question that attempts to seek
boundaries: does art have to be made by an artist? If art can be defined
by an audience's reception of an object or event and not by a performer (e.g.
artist-- regardless of intent), then a micrograph can be a work of art
without there being an artist, in this case a scientist. The danger this
position raises, however, is that it may not priviledge only the relics of
human endeavor; for instance, a rock or sunrise could be misconstrued as art.
Here is the center of the issue: the objects to which the questions really
refer (micrographs) are compelling aesthetic objects. Just because
something looks good or touches that ineffable something in us and is made
by a person or people does not mean it is art.

None of these musings lead directly to the question of why there
are micrograph competitions. Entering a competition implies desire for
exhibition or need for public recognition or motivation to WIN or the
expectation of another line on the cv, but these
apects of showing work, although the same or similar to concerns in the
arts, really have nothing to do with art specifically.

The only way, logically, to answer the general question whether scientists are
artists and to, pragmatically, avoid exploding the term "artist" so that
it is meaningless, is to answer NO. More simply: There are scientists who
are artists. Not all scientists are artists. The last two statements
have nothing to do with whether micrographs, in general, are art. Some
micrographs are art and some are not.

-mc






From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Thu, 23 Jun 1994 17:41:46 -0400 (EDT)
Subject: Re: Is EM art?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

EM Bulletin Board {microscopy-at-anlemc.msd.anl.gov}
In-Reply-To: {Pine.3.89.9406231757.A17083-0100000-at-umbc8.umbc.edu}
Message-Id: {Pine.3.07.9406231745.A6730-9100000-at-alsys1}
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; charset=US-ASCII

On Thu, 23 Jun 1994, rutledge phil wrote:
} On 23 Jun 1994, Paul Webster wrote:
} } OFFICE MEMO Is EM art? Date: 6/23/94
} } Are we scientists are artists? If the answer is the first then why are there
} } so many micrograph competitions?
} } To answer this.
} EM is an art! You either have it or you don't!
EM is a craft.






From: tayloe-at-rorc.usbm.gov
Date: Thu, 23 Jun 1994 16:50:56 -0500 (CDT)
Subject: Re: Is EM art?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On 23 Jun 1994, Paul Webster wrote:

} Subject: Time: 4:42 PM
} OFFICE MEMO Is EM art? Date: 6/23/94
} Are we scientists are artists? If the answer is the first then why are there
} so many micrograph competitions?

From the for what its worth dept., since ya asked...:

I don't think that being a scientist should exclude one from appreciating
art. Hopefully, its in the study of the symmetry and beauty of nature
(be it biological or material, with a light or electron 'scope) that
scientists can learn and "see" more of the world around them. Having such
competitions, to myself anyway, allows one to view others' work; it gives
a forum for the display of different photomicros that otherwise I might
not ever see for myself. It also allows others to see the best of
different techniques/methods.

Besides, its a way to win neato prizes :) :)
But I suppose that ya have to enter 'em first, 'eh?

My {$0.02,
-Rob
,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,
| Rob Tayloe | MSM Spelunkers Club /\v/\ |
| Metallographic Lab. | Missouri Speleological Survey \-v-/ |
| Rolla Research Center | Bat Conservation International /\v/\ |
| U.S. Bureau of Mines | Missouri Cave & Karst Conservancy |
| tayloe-at-rorc.usbm.gov | National Speleological Society #32993 /-v-\ |
| (314) 364-3169 x247 | American Cave Conservation Association |
''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''





From: George M :      George_M-at-image1.com
Date: Thu, 23 Jun 1994 15:04:29 -0700
Subject: What's Appropriate? Please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


From HOLONET What's Appropriate? Please stick to microscopes!
Reply to Tobias Baskin's message:

I was away from my office for one week and came back to over 150 E-mail
messages from the Confocal newsgroup and from this newsgroup. It's going to
take a long time to read and save the useful messages, and delete the redundant
replies, and delete the all too large number of off-topic messages. Come to
think of it, this reply is probably off-topic, so I would aprreciate your
re-reading the Moderator's announcements of the purpose of this newsgroup
instead of replying to this message.

} I would like to say that the discussion on overstriking has been very
helpful...

I deleted all of those with extreme prejudice without reading any.

} I bet word processing is the ONLY technique that EVERYone on this list
uses....

So what. I'm on here to learn about MICROSCOPES. I'll go as far as
anti-vibration tables for acceptable hardware. Questions and comments on
software not directly related to to your microscope should be asked elsewhere.

} there are many subjects covered on the list that I ignore...

Yes, but the junk mail still shows up in my mailbox.

} Three cheers for clear subject lines.

Agreed. I will delete without reading all future messages that have "None" or
garbled subject lines (my apologies to any customers who don't get a reply from
me because of this policy).


Sincerely,
George McNamara








From: P.V.Hatton-at-sheffield.ac.uk
Date: 23 Jun 94 18:47:32
Subject: Microscopy, not word-processors!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.940624083209.352-at-bunyip.ph.rmit.edu.au}
To: microscopy-at-anlemc.msd.anl.gov

I agree entirely
John Millar RMIT Australia
------- Forwarded Message Follows -------
To: microscopy-at-anlemc.msd.anl.gov

Hi, I would have kept quiet but for some users suggesting that it is
a good idea to exchange trivia about word processing. I joined a
microscopy board, not a software panel - I get a lot of useless mail
and I don't need a pile of messages about word processing. May I
politely suggest that people who want to talk about anything except
microscopy on this valuable service should form their own club.

Votes for and against?

Paul Hatton
University of Sheffield






From: Paul Webster :      Paul_Webster-at-quickmail.yale.edu
Date: 23 Jun 1994 18:40:58 -0400
Subject: Re: is EM art?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


bla bla_ Re} is EM art?
If we are artists, why do we submit our work to scientific journals?






From: ldm-at-apollo.numis.nwu.edu (L. D. Marks)
Date: Fri, 24 Jun 1994 08:27:32 -0500
Subject: HREM simulation software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Paul Fons asked about HREM image simulation software. In this
context I would like to float an idea for responses. In the past,
Hitachi has been selling our software at prices which (in my opinion)
are ludicrously high. We are in the process of rewriting it for a
RISC/UNIX workstation, exploiting X-windows and UNIX wide system calls.
I am very tempted to make it freeware, either via ftpanon or otherwise
when this is finished; in other words it should be compilable on
any modern UNIX system (and PC/Mac systems as they move to be
compatible with workstations).

Comments ?




From: A. Kent Christensen :      akc-at-umich.edu
Date: Fri, 24 Jun 1994 09:48:29 -0400 (EDT)
Subject: Re: Is EM art?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Paul, old man,

As you know, it is a fallacy to imply that the two are mutually
exclusive. EM is used as a tool in a great many areas of science. The
yearning to use such a tool with some elegance (rather than in a manner
that is dull or sloppy) is not unique to EM. I know biochemists,
molecular biologists, and others, whose work has elegance. They are not
satisfied with a mere dumping of required information (in any form) into
a paper. They want the information to be well organized, tight, clear, a
pleasure to read. One might say they do science with some art. I hope
we can have more science involving EM that has this kind of elegance. If
there is an occasional EM art show on the side, it is certainly not a
problem.

Regards,
Kent
(A. Kent Christensen, Dept. of Anatomy and Cell Biology, Univ of
Michigan, {akc-at-umich.edu} )

-----------------------------------

On 23 Jun 1994, Paul Webster wrote:

} Subject: Time: 4:42 PM
} OFFICE MEMO Is EM art? Date: 6/23/94
} Are we scientists or artists? If the answer is the first then why
are there } so many micrograph competitions?
}
}
}




From: ldm-at-apollo.numis.nwu.edu (L. D. Marks)
Date: Fri, 24 Jun 1994 08:27:32 -0500
Subject: HREM simulation software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reply_ RE} HREM simulation software
Reply
from:
John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX: (313)936-3352
jfmjfm-at-umich.edu or John_Mansfield-at-mse.engin.umich.edu
Absolutely excellent idea!!!!!
There are too many people who write software in the microscopy field and think
they are going to get rich on it. They try to sell it for exorbitant amounts of
money and it is frequently buggy as all hell.They are usually unable to support
it adequately since they are typically only selling it as a part time job.
Software marketing and support is difficult to do properly!
I think making it available free would be an excellent service to the user
community.

--------------------------------------
Paul Fons asked about HREM image simulation software. In this
context I would like to float an idea for responses. In the past,
Hitachi has been selling our software at prices which (in my opinion)
are ludicrously high. We are in the process of rewriting it for a
RISC/UNIX workstation, exploiting X-windows and UNIX wide system calls.
I am very tempted to make it freeware, either via ftpanon or otherwise
when this is finished; in other words it should be compilable on
any modern UNIX system (and PC/Mac systems as they move to be
compatible with workstations).

Comments ?

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;24 Jun 1994 09:48:06 U
Received: by totalrecall.rs.itd.umich.edu (8.6.9/2.3)
with X.500 id JAA22539; Fri, 24 Jun 1994 09:47:48 -0400
Received: from anlemc.msd.anl.gov by totalrecall.rs.itd.umich.edu (8.6.9/2.3)
with SMTP id JAA22527; Fri, 24 Jun 1994 09:47:45 -0400








From: John Posthill :      jbp%rtifs2-at-rti.rti.org
Date: Fri, 24 Jun 1994 10:44:23 -0400 (EDT)
Subject: Re: HREM simulation software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Why not? How can anyone be against this? If you can cover your costs
somehow, and you want to do it, consider me a cheerleader!

On Fri, 24 Jun 1994, L. D. Marks wrote:

} Paul Fons asked about HREM image simulation software. In this
} context I would like to float an idea for responses. In the past,
} Hitachi has been selling our software at prices which (in my opinion)
} are ludicrously high. We are in the process of rewriting it for a
} RISC/UNIX workstation, exploiting X-windows and UNIX wide system calls.
} I am very tempted to make it freeware, either via ftpanon or otherwise
} when this is finished; in other words it should be compilable on
} any modern UNIX system (and PC/Mac systems as they move to be
} compatible with workstations).
}
} Comments ?




From: jmcgee-at-lunatic.er.usgs.gov (Jim McGee (703)648-6782)
Date: Fri, 24 Jun 1994 15:51:17 -0400
Subject: Ni standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

standard-at-lunatic.er.usgs.gov


A synthetic Ni-olivine (Ni2SiO4) from the Smithsonian (USNM #717) may be
available. We have some of this material that was used in a thermodynamic
study by R. Robie et al (Amer. Mineralogist v.69, p.1096, 1984). You may
be able to obtain some through Gene Jarosewich at the Smithsonian.

Jim McGee


James J. McGee email: jmcgee-at-lunatic.er.usgs.gov
U.S. Geological Survey Phone: (703) 648-6782
959 National Center Fax: (703) 648-6789
Reston, VA 22092





From: tivol-at-tethys.ph.albany.edu
Date: Fri, 24 Jun 1994 16:45:43 EDT
Subject: Nickel standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear John,
I assume the unknowns are silicate minerals. If not, you should make
an attempt to have the same average Z-value for unknowns and standards--Chuck
Fiori made a point of this in referrence to biological specimens. I'd try to
mix a nickel compound (any one which is handy and doesn't have other peaks in
the energy region of interest will do) with a molten silicate glass or with
quartz if you can heat it enough to get thorough mixing. I'd make a range of
different nickel fractions from ~1% to whatever you expect at maximum. The
average Z and thorough mixing are the keys. Good luck.

Yours,

Bill Tivol




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Fri, 24 Jun 1994 16:53:38 -0400 (EDT)
Subject: Suitable subjects for discussion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



As long as we are all voting, I agree that the discussion of word
processors has run its course. However I did learn some useful things
applicable not just to miller indexes but to microscopy writing in
general. Since we are a varied background of users (and many of us could
care less about subscribing to a bulletin board on word processors or
operating systems, etc.) . IMHO the occasional inclusion of helpful but
peripheral information and questions should be encouraged. Like others,
my mailbox is too cluttered and I throw out a lot without reading. But I
would like to keep things open. Occasionally a gem comes along.

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************





From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sun, 26 Jun 1994 21:08:05 -0500 (CDT)
Subject: LM - microscope rotating stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English




From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sun, 26 Jun 1994 21:08:05 -0500 (CDT)
Subject: LM - microscope rotating stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


RFC-822-Headers:
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; charset=US-ASCII
On Tue, 21 Jun 1994, Carlo Montemagno wrote:

} I need to purchase an inverted epi microscope to conduct membrane kinetic
} studies. Because of the experimental setup I can't use immersion optics. I have
} about 22K to spend and I already have a good quality video camera. Are there
} any thoughts on which product lines are the best and recomentations for
} objectives?
}
} Thanks
}
} Carlo Montemagno

On June 22nd Clarissa wrote in response to Carlo Montemagno's question:

} Try a Nikon, if you want a rotating stage a Nikon will not do.

For those who want a rotating stage for the Nikon inverted microscope,
as of a few weeks ago, a rotating, centrable, gliding stage is available for
the Nikon Diaphot 200/300.
It attaches by the same 3 screws as the standard stages, it is not graduated,
but works well for DIC. Cost: $1600.00
It is manufactured by Micro Video Instruments of Avon, Ma. USA
For more information: 508 580 0080 or FAX 508 580 8623

Erwin Deutsch
MVI
IN%"/PN=erwin.deutsch/O=MVI/ADMD=TELEMAIL/C=US/-at-sprint.com"










From: andy-at-earwax.pd.uwa.oz.au (Andy Johnson)
Date: Mon, 27 Jun 1994 10:39:12 +0800
Subject: Re: HREM simulation software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I also heartily agree that open software would be a big advance. Don't
limit it to HREM. The errors in commercial packages are virtually
uncorrectable and we wind up with the situation bulldust in (the code) =
bulldust out. Sadly only known after a lot of pain and wasted time. eg the
K-line simulation problem tracked down by Alwyn Eades and his
collaborators.

Andy Johnson, CMM, UWA







From: Nestor J. Zaluzec-ANL EMCenter :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Sun, 26 Jun 1994 22:23:05 -0500 (CDT)
Subject: Imaging Programs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In another "hot" topic which could set off a mine is better than
yours email war.....

JEC of GIT asked about imaging programs (and possible platform
competition). Well let me try to be impartial and start the
discussion ;-)
------------------------------------------------------------

There are many commerical programs available for handling
image data, all of these do good to excellent jobs, (on both
Mac's and PC's -- Oops just showed a preference didn't I--)
unfortunately you will have to purchase them.

But to answer your question .
No one has bothered to port NIH image to the PC platform. The
lastest version 1.56beta9 runs on virtually all Mac's but not the
PC. The source code is available on Zippy.nimh.nih.gov if you
are interested in a challenge.

On the PC side the only freeware programs I am familiar with
are NCSA Image for the PC, and UCFImage for the PC. There
are also a couple of programs which I would call Viewers, the
just let you put images up on the screen and not much else.
All of these are freeware and will be available at the computer
workshops at the PARIS ICEM13 meeting (July 17-22), as well as the
at the New Orleans MSA/MAS meeting (August 1-5).


The NCSA suite of programs were not designed as an image processing
programs but then again neither was NIH-Image, rather they might be
better described as a data set visualization programs. I've
occassionally used the NCSA code but not frequently, since I've found
that I can more readily customize NIH Image to Microscopy
related tasks.

UCFImage is a older image processsing type program out of the
University of Central Florida, the version (5.0) I have lacks
the polish of a true GUI program (it runs under DOS)
and hence to most users who have been wedded to WINDOWS or Mac's
(there I've now balanced my comments) might be disappointed.
I found it clumbersome to use, however, if you are running on
an older DOS machine then you may wish to check it out.


If anyone is still interested I can dig up the readme files
and post segments tomorrow.

I'd be interesting in hearing about any other public domain
programs which can be added to the software library database.
So if you know of such please post the information here.
as well as include the information on how to get copies.
I'll then add the information to the EMMPDL libraries and
try to insure that copies are available next month at both
ICEM and MSA.

Nestor J. Zaluzec
ANL






From: andy-at-earwax.pd.uwa.oz.au (Andy Johnson)
Date: Mon, 27 Jun 1994 11:47:40 +0800
Subject: Re: HREM simulation software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Received: From CORNELLC(MAL) by ANLNRH(ANJE6.10) for MICROSCO-at-ANLEMC; Sun, 26
Jun 94 22:49
Return-Path: {-at-CORNELLC.CIT.CORNELL.EDU:andy-at-EARWAX.PD.UWA.OZ.AU}
Received: from CORNELLC (NJE origin SMT-at-CORNELLC) by CORNELLC.CIT.CORNELL.EDU
(LMail V1.1d/1.7f) with BSMTP id 0700; Sun, 26 Jun 1994 23:47:52 -0400
Received: from uniwa.uwa.edu.au by CORNELLC.cit.cornell.edu (IBM VM SMTP V2R2)
with TCP; Sun, 26 Jun 94 23:47:49 EDT
Received: from earwax.pd.uwa.edu.au (earwax.pd.uwa.edu.au [130.95.156.3]) by
uniwa.uwa.edu.au (8.6.8/8.6.4) with ESMTP id LAA22954 for
{Microscopy-at-anlemc.BITNET} ; Mon, 27 Jun 1994 11:48:46 +0800
Received: from [130.95.124.120] (awsj [130.95.124.120]) by earwax.pd.uwa.edu.au
(8.1C/8.1) with SMTP id LAA10562; Mon, 27 Jun 1994 11:47:40 +0800

I also heartily agree that open software would be a big advance. Don't
limit it to HREM. The errors in commercial packages are virtually
uncorrectable and we wind up with the situation bulldust in (the code) =
bulldust out. Sadly only known after a lot of pain and wasted time. eg the
K-line simulation problem tracked down by Alwyn Eades and his
collaborators.

Andy Johnson, CMM, UWA







From: andy-at-earwax.pd.uwa.oz.au (Andy Johnson)
Date: Mon, 27 Jun 1994 13:30:40 +0800
Subject: Re: HREM simulation software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Received: From ANLVM(MAILER) by ANLNRH(ANJE6.10) for MICROSCO-at-ANLEMC; Mon, 27
Jun 94 00:32
Received: from ANLVM by ANLVM (Mailer R2.07B) with BSMTP id 3970; Mon, 27 Jun
94 00:31:23 CDT
Received: from earwax.pd.uwa.edu.au by ANLVM.CTD.ANL.GOV (IBM VM SMTP
R1.2.2ANL-MX) with TCP; Mon, 27 Jun 94 00:31:19 CDT
Received: from [130.95.124.120] (awsj [130.95.124.120]) by earwax.pd.uwa.edu.au
(8.1C/8.1) with SMTP id NAA11095; Mon, 27 Jun 1994 13:30:40 +0800

I also heartily agree that open software would be a big advance. Don't
limit it to HREM. The errors in commercial packages are virtually
uncorrectable and we wind up with the situation bulldust in (the code) =
bulldust out. Sadly only known after a lot of pain and wasted time. eg the
K-line simulation problem tracked down by Alwyn Eades and his
collaborators.

Andy Johnson, CMM, UWA







From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Mon, 27 Jun 1994 08:33:50 +0000 (GMT)
Subject: EDX: Kevex u-7000 part wanted

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

Anyone have a K-1905-1 board out of a Kevex micro-7000 EDX analyser spare
that they are able to give away / sell? Ours is faulty and hindering one of
our M.Sc. students.

Thankyou,

Keith Hallam

p.s. If I send a message out on the microscopy discussion group does it
automatically get echoed to the usenet group too? If so, apologies for repeating
myself!







From: NAME :      ZALUZEC-at-anlemc.msd.anl.gov
Date: Mon, 27 Jun 1994 20:50:54 -0500 (CDT)
Subject: Network Problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



G'day Subscribers:

We had a major network crash here at ANL over the weekend.
FRIED fiber optics and lots of headaches all around. I
apologize if messages got lost over the weekend. I think
things are back to "semi-normal".

Nestor
Your friendly neighborhood SysOp...




From: Antonio Jorge Silvestre :      fajs-at-skull.cc.fc.ul.pt
Date: Tue, 28 Jun 1994 12:01:48 +0000
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: fajs-at-scosysv.cc.fc.ul.pt

unsubscribe





From: andy-at-earwax.pd.uwa.oz.au (Andy Johnson)
Date: Tue, 28 Jun 1994 16:28:06 +0800
Subject: Re: HREM simulation software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Received: From ANLVM(MAILER) by ANLNRH(ANJE6.10) for MICROSCO-at-ANLEMC; Tue, 28
Jun 94 03:29
Received: from ANLVM by ANLVM (Mailer R2.07B) with BSMTP id 7712; Tue, 28 Jun
94 03:28:47 CDT
Received: from earwax.pd.uwa.edu.au by ANLVM.CTD.ANL.GOV (IBM VM SMTP
R1.2.2ANL-MX) with TCP; Tue, 28 Jun 94 03:28:44 CDT
Received: from [130.95.124.120] (awsj [130.95.124.120]) by earwax.pd.uwa.edu.au
(8.1C/8.1) with SMTP id QAA17300; Tue, 28 Jun 1994 16:28:06 +0800

I also heartily agree that open software would be a big advance. Don't
limit it to HREM. The errors in commercial packages are virtually
uncorrectable and we wind up with the situation bulldust in (the code) =
bulldust out. Sadly only known after a lot of pain and wasted time. eg the
K-line simulation problem tracked down by Alwyn Eades and his
collaborators.

Andy Johnson, CMM, UWA







From: Paul C. Abbott :      paul-at-earwax.pd.uwa.edu.au
Date: Wed, 29 Jun 1994 10:51:02 +0800
Subject: test

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


test




From: PEARSON-at-ANMSD3.MSD.ANL.GOV (JOHN E. PEARSON 708.252.7738, 9595(FAX))
Date: Wed, 29 Jun 1994 9:09:33 CDT
Subject: W crucible source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a vendor for tungsten (W) crucibles? I am looking for
a 6mm diameter W tube closed one end.




From: randy nessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Tue, 28 Jun 1994 15:32:22 -0500 (CDT)
Subject: Diffraction pattern software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



The image analysis facility down the hall is interested in the names of
some software packages that a digitized or scanned electron diff pattern can be
measured and indexed with. Are there any public domain/shareware programs
available? Since the demand is small, they aren't willing to spend a great
deal. Thanks in advance.
Randy

___________________________________________________________________
Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu
The University of Iowa Central Electron Microscopy Research Facility






From: Paul C. Abbott :      paul-at-earwax.pd.uwa.edu.au
Date: Wed, 29 Jun 1994 10:53:38 +0800
Subject: test

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Received: From CORNELLC(MAL) by ANLNRH(ANJE6.10) for MICROSCO-at-ANLEMC; Tue, 28
Jun 94 21:55
Return-Path: {-at-CORNELLC.CIT.CORNELL.EDU:paul-at-EARWAX.PD.UWA.EDU.AU}
Received: from CORNELLC (NJE origin SMT-at-CORNELLC) by CORNELLC.CIT.CORNELL.EDU
(LMail V1.1d/1.7f) with BSMTP id 2085; Tue, 28 Jun 1994 22:53:48 -0400
Received: from uniwa.uwa.edu.au by CORNELLC.cit.cornell.edu (IBM VM SMTP V2R2)
with TCP; Tue, 28 Jun 94 22:53:46 EDT
Received: from earwax.pd.uwa.edu.au (earwax.pd.uwa.edu.au [130.95.156.3]) by
uniwa.uwa.edu.au (8.6.8/8.6.4) with ESMTP id KAA04414 for
{Microscop-at-anlemc.BITNET} ; Wed, 29 Jun 1994 10:54:44 +0800
Received: from localhost (paul-at-localhost) by earwax.pd.uwa.edu.au (8.1C/8.1) id
KAA19840; Wed, 29 Jun 1994 10:53:38 +0800

test




From: Marcelle A Gillott :      magem-at-csd4.csd.uwm.edu
Date: Thu, 30 Jun 1994 09:32:04 -0500 (CDT)
Subject: em-art

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Why can't EM be both science and art - we use this tool as scientist and
in the process create some wonderfully beautiful images that have a
definite aesthetic value - in my opinion the fact that we can appreciate
the beauty and have a desire to share it in no way detracts from our
seriousness as scientists - besides it may grab the attention and
interest of a non-scientist and get him or her interested in learning
more about the subject of the image or the process by which it was
created

Marcelle Gillott
UWM






From: rms-at-vax.ox.ac.uk
Date: Thu, 30 Jun 1994 14:31:33 +0100
Subject: July issue of the Journal of Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk
74250.331-at-COMPUSERVE.COM, ROSS.MACKENZIE-at-OX.AC.UK,
MICROSCOPY-at-ANLEMC.MSD.ANL.GOV
Message-ID: {00980BAD.310E5BC9.123-at-vax.ox.ac.uk}

JULY 1994 ISSUE OF THE JOURNAL OF MICROSCOPY - VOLUME 175 PART 1

Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 1-9

Differential imaging in confocal microscopy

T. WILSON, R. JUSKAITIS & J. B. TAN, Department of Engineering, University of Oxford,
Parks Road, Oxford, OX1 3QZ, U.K.


SUMMARY
A coherent detection system using a two-mode optical fibre is described which permits
confocal, differential amplitude contrast and differential phase contrast images to be obtained
simultaneously from a scanning optical microscope. the direction of differentiation may be
chosen arbitrarily. The differential imaging modes possess an optical sectioning property.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 10-20

A novel method of Z-contrast imaging in STEM applied to double labelling

W. TICHELAAR, C. FERGUSON, J.-C. OLIVO, K. R. LEONARD & M. HAIDER,
European Molecular Biology Laboratory, Meyerhofstrasse 1, D-69012 Heidelberg, Germany


SUMMARY
A novel method of Z-contrast imaging in the scanning transmission electron microscope
(STEM) is presented. The technique relies on the element dependence of the angular
distribution of the scattered electrons, and is realized with a detector consisting of a set of
concentric rings. It is possible to discriminate 9-nm colloidal gold and silver specifically
distributed on thin sections. In addition to this practical work, numerical evaluations are used
to assess the method. With two smaller markers, this approach will be useful in discriminating
closely-spaced antigenic sites when steric hindrance occurs with double-labelling using
probes of different sizes.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 21-33

A new method to correlate acoustic spectroscopic microscopy (30MHz) and light
microscopy

A. F. W. VAN DER STEEN, J. M. THIJSSEN, J. A. W. M. VAN DER LAAK, G. P. J.
EBBEN & P. C. M. DE WILDE, University Hospital Nijmegen, PO Box 9101, 6500 HB
Nijmegen, The Netherlands


SUMMARY
A powerful new method is used to correlate between light microscopic and acoustic properties
of biological tissues. Specimens of liver were sectioned into slices, between 250 and 10
micrometres thick. The thick sections were investigated acoustically, the thin sections by light
microscopy. Markers that could be detected and located, both optically and acoustically, were
used to find and reconstruct corresponding regions in the acoustic and optical sections (2.5
x 2.5 mm).
Parameter images were reconstructed from the sections investigated acoustically. The
acoustic parameters were attenuation at 30MHz, the slope of the attenuation spectrum
(between 10 and 50MHz), backscattering at 30MHz, the slope of the backscattering spectrum
(between 10 and 50MHz), and the local ultrasound velocity. Acoustic images were obtained
in the frequency range from 10 to 50 MHz, yielding a lateral resolution of about 50
micrometres.
The sections for light microscopy were stained according to the Goldner trichrome
staining technique. The histological composition was determined quantitatively, using digital
image segmentation techniques. The percentage of collagen-rich fibrous tissue, luminal
structure and interstitial spaces, and the number of nuclei were calculated for regions of 250
x 250 micrometres. These histological features were correlated with acoustic parameters
obtained from the corresponding regions in adjacent sections. It was thus possible to find the
histological components responsible for acoustic parameters.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 34-43

High pressure freezing of cell suspensions in cellulose capillary tubes

H. HOHENBERG, K. MANNWEILER & M. MULLER*, *Laboratory for Electron
Microscopy I, Institute of Cell Biology, Universitatsstrasse 2, CH-8092 Zurich, Switzerland.
and Heinrich-Pette-Institute for Experimental Virology and Immunology, at the University
of Hamburg, D-20251 Hamburg, Germany


SUMMARY
A procedure for efficient cryoimmobilization of large volumes of cell suspensions or micro-
organisms by high-pressure freezing is described. This procedure uses transparent, porous
cellulose capillary tubes with an inner diameter of 200 micrometres, into which the
suspensions are drawn by capillary action. The tubes are processed by high-pressure freezing
and freeze substitution as if they were tissue samples. Centrifugation of suspensions at low
temperatures is no longer necessary and cryopreparation is greatly facilitated.
A very high yield of adequately frozen specimens is obtained due to the constant,
defined sample geometry. This approach can also be used to process suspensions by
conventional chemical fixation, eliminating the need to embed pellets in low-melting-point
agarose, for example, prior to chemical fixation.
The preparation procedure is demonstrated with suspensions of nematodes, paramecia
and bacteria.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 44-53

Nucleotide and protein distribution in BrdU-labelled polytene chromosomes revealed by
ion probe mass spectrometry

RICCARDO LEVI-SETTI, JAN M CHABALA & SARAH SMOLIK*, Enrico Fermi
Institute, The University of Chicago, 5640 South Ellis Avenue, Chicago IL 60637, U.S.A.,
and *Vollum Institute for Advanced Medical Research, Oregon Health Services University,
Portland, OR 97201, U.S.A.


SUMMARY
Detailed chemical maps of BrdU-labelled polytene chromosomes of Drosophila melanogaster,
obtained by imaging secondary ion mass spectrometry, reveal separately the distribution of
DNA and proteins in the chromosomes. The thymidine-analogue BrdU within the
chromosomal DNA is localized by detecting the Br- secondary ion signal, while both nucleic
acid and protein content, are mapped through the abundantly emitted CN- signal. This novel
approach supersedes, and helps to explain the origin of, the banding patterns that are observed
by conventional staining techniques. the high spatial resolution and chemical and isotropic
sensitivity of the technique should enhance the localization of specific genes by in situ
hybridization in mitotic chromosomes.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 54-59

Application of backscattered electron imaging to the study of the contact zone between
lichen and rock. Assessment preparative procedures

J. WIERZCHOS & C. ASCASO, Centro de Ciencias Medioambientales, Serrano 115, bis
Madrid 28006, Spain


SUMMARY
In the study of the lichen-rock interface, light microscopy, scanning electron microscopy
(SEM) in secondary emission mode and transmission electron microscopy are the most
commonly-used techniques. As these methods have some limitations, there is a need to
explore other techniques for observation of the lichen-substrate interface. One of the most
promising methods is the application of SEM in the back-scattered electron (BSE) emission
mode.
The thallus of Aspicilia intermutans (Nyl.) Arn. growing on granitic rock was
examined by SEM in BSE mode. The detailed preparation of transverse sections of the
lichen-rock contact zone is presented. The BSE scanning images of the lichen-rock interface
obtained present new insights into the ultrastructural features of the biological components,
providing more information about the biogeophysical and biogeochemical weathering of rock.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 60-69

Semi-automated measurement of true chord length distributions and moments by video
microscopy and image analysis

EBEN H. OLDMIXON, JAMES P. BUTLER* & FREDERIC G. HOPPIN, Division of
Pulmonary Medicine, Memorial Hospital of Rhode Island, 111 Brewster Street, Pawtucket,
RI 02860, U.S.A. and *Harvard School of Public Health, Boston MA 02115-9957, U.S.A.


SUMMARY
The distribution of lengths of airspace chords in pulmonary parenchyma characterizes many
architectural features of the alveoli and alveolar ducts. Laborious to obtain manually, the
distributions and density functions may be acquired semi-automatically by video microscopy,
digitization and image processing. The accuracy of the estimation is influenced by the
microscopical methods and also by the techniques used (i) to convert the digitized grey-scale
picture to a two-valued image, (ii) to collect the chord lengths and (iii) to compensate for
finite field widths. The last problem arises because some of the chords are completely visible
within a field, while others are only partially seen, since one of the two air-space boundaries
lies outside the field of view. The error systematically biases the observed distribution. This
paper presents solutions to hardware, software and analytic problems encountered while
developing the capability to measure airspace chord length density functions semi-
automatically. Formulas for estimating the true chord length density function from samples
of observed chord lengths are presented. Also given are formulas for the estimation of the
first and second moments of the true chord length distribution from the means of observed
chord lengths. These techniques of image preparation and analysis should be suitable for
characterizing particle, grain or cell size distributions, especially where many profiles fall
partially outside the field of view.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 70-83

A simple algorithm to measure the volume-weighted and number-weighted mean
volume of particles

GUY J. LAROYE & K. GRANT, Ontario Cancer Institute, 500 Sherbourne Street, Toronto,
Ontario M4X 1K9, Canada


SUMMARY
An algorithm is presented which offers an alternative approach for measuring volume- and
number-weighted mean volume and standard deviation of particles. Using a computer-
assisted manual method the following intermediate steps are performed automatically:
generation of linear probes emanating from the sampling point of the object and intersecting
the profile periphery, measurement of their lengths, and measurement of the area of the
transect required for estimating the standard deviation of the volume-weighted mean volume.
By first tracing manually the outline of the periphery of the object with a cursor, on a
magnetic tablet or on an image acquired into the computer with a video camera, the location
of all pixels of the periphery is registered and the area of the transect is measured
concurrently. The computer is informed of the coordinates of the selection point in the
uniform random (UR) sampling grid by clicking the cursor. All ensuing random linear probes
between the sampling point and the object profile periphery emanating from this selection
point, radiating at angular intervals of 29-30 degrees to the periphery. In the case of vertical
sections, similar lines are generated at intervals where the sine of the angle changes by a
value of 0.33. The volume-weighted mean volume of the object is estimated from the average
of all products. As the periphery is traced, the algorithm can automatically determine the area
of the cross-section of the object, from which the standard deviation of the volume-weighted
mean volume can be calculated. Some elements of the algorithm are also used for the
measurement of the number-weighted mean volume. The latter procedure is facilitated using
an acoustic vertical depth monitor attached to the microscope. The impact of truncation ('lost
caps') on the precision of measurements is discussed. The algorithm is of particular use in
light microscopy for measuring cell nuclei by direct visual inspection of the microscopic field
using a side-arm mirror assembly interfaced with a magnetic tablet.



Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 84-89

Microspectroscopic measurement of optical properties of rat liver in the visible region

AKITOSHI SEIYAMA, SHENG-SONG CHEN, HIROAKI KOSAKA & TAKESHI SHIGA,
Department of Physiology, Osaka University Medical School, 2-2 Yamadaoka, Suita, Osaka
565, Japan


SUMMARY
Microspectroscopy is used to investigate optical properties of haemoglobin-free perfused rat
liver. Visible spectra of 20 micrometre diameter spot size were measured in transmission
and/or reflection modes as a function of the thickness ( {1200 micrometres) of the liver edge.
Optical density (OD) in transmission mode increased with increasing liver thickness, whereas
in reflection mode OD decreased but became almost constant above a certain thickness (c.600
micrometres) of the liver. The Kubelka-Munk (KM) two-flux model, with a minor
modification, was applied successfully to the analysis of the changes in OD as a function of
the thickness. This approach estimates the KM absorption coefficient, KM scattering
coefficient and effective penetration depth of the liver. The optical properties were similar to
reported values, obtained with different methods.


Journal of Microscopy, Vol. 175, Pt. 1, July 1994, pp. 90

Letter to the Editors: A generalized terminology for multidimensional microscopy

MARK A BROWNE, C VYVYAN HOWARD, GLEN JOLLEYS & DUNCAN STACEY,
Department of Fetal & Infant Pathology, University of Liverpool, PO Box 147, Liverpool,
L69 3BX,






************************************************************************

Gillian Wilson Executive Editor
The Royal Microscopical Society Journal of Microscopy &
37/38 St Clements Proceedings of the RMS
Oxford
OX4 1AJ UNITED KINGDOM
rms-at-uk.ac.ox.vax OR rms-at-vax.ox.ac.uk
Telephone +44 (0)865 248768
Fax +44 (0)865 791237

************************************************************************




From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Thu, 30 Jun 1994 11:19:24 -0700
Subject: Re: ultramicrotomes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Comments: Converted from PROFS to RFC822 format by PUMP V2.2X

} ALSO...we have an OMU-3 which we LIKE to repair, but I was told some time
} ago that parts for it are also no longer available. Is anyone aware of
} a parts inventory and/or service for this instrument.
}

Try calling TEK-NET, Inc. I haven't used them, but spoke with someone
there recently, so I'm pretty sure they're still in business. Their letter
of introduction, sent out in '90, described the organization as a group of
service engineers who used to work at Cambridge Instruments' East Region
Service Center, until that center was closed. The list of equipment that
they service is impressive, and Reichert is among them.

I've been tempted to post this information before, just to let folks know
they exist, so here it is:

TEK-NET, Inc.
1985 Swarthmore Ave.
Lakewood, New Jersey 08701
(908) 905-5530
(908) 905-0152 FAX

I have no affiliation with them, just wanted to get the word out.


John chandler-at-lamar.ColoState.EDU Fort Collins CO






From: EMLAB-at-opus.mco.edu
Date: Thu, 30 Jun 1994 09:55:28 -0400 (EDT)
Subject: LM TEXTBOOK

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This fall we are starting a new graduate level course on all aspects
of light microscopy ie. fluorscence(sp),polorized,bright/dark field etc. and
need recommendations for a good textbook. This course will include epquiment
and methods. Thanks in advance for any help.

Ed Calomeni




From: Alan Wilson :      wilsona-at-blackjack.cis.dsto.gov.au
Date: Fri, 1 Jul 94 10:26:45 EST
Subject: N2 Vacuum Cleaning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





MicroscopyListserver Archive Email Extraction Software Version NJZ07060908

Return to Microscopy Listserver Home Page


Return to MSA HomePage