From: {magem-at-csd.uwm.edu}:ddn:wpafb
Date: 9-30-94 9:59am
Subject: Re: magnetic SEM specimen

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: magnetic SEM specimen
Orig-Author: {Marcelle A Gillott {magem-at-csd.uwm.edu} }:ddn:wpafb
-----------------------------------------------------------

help!

I am a biologist with limited experience with materials specimens.......
a student wants to look at a magnetic sample and I would appreciate
any suggestions as to the best way to handle this?

I am using a s570 (Hitachi)

thanks in advance

marcelle






From: JJMILL-at-bunyip.ph.rmit.oz.au
Date: Tue, 4 Oct 1994 08:48:56 EST-10
Subject: (Fwd) LM: Source for 49 C paraffin?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941004084856.448-at-bunyip.ph.rmit.edu.au}
To: microscopy-at-aaem.amc.anl.gov

on 4 October, Dwight Beebe wrote :

Wtih respect to the "kill the feed" demand, I personally find the
discussion of other techniques and aspects of EM/LM to be
interesting and
often directly useful in my research and teaching. I want to thank
Nestor for his continued work and all the others who contribute to the
discussions for making this a very productive source of information.

.......I would like to agree that the list serves a very useful function,
even though one accumulates a lot of messages during an absence,
as I did recently. However, the availability of the wealth of expertise
at one's fingertips is invaluable. The unhelpful and silly comments
only detract from its value, and if people would take the time to read
the information that is available, many of the objections would not be
necessary. My thanks also to Nestor and all of the contributors.

The person looking for Super SEM from Griffin and Van Riessen could
try Arie at avr-at-uniwa.uwa.edu.au
cheers
jjm
Professor John J. Millar
Head, Department of Applied Physics
Electron Microscope Unit
RMIT, GPO Box 2476V, MELBOURNE AUSTRALIA 3001
+613 660 2602 fax 613 660 3837
email jjmill-at-rmit.edu.au




From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Tue, 4 Oct 1994 11:01:02 -0500
Subject: Methyl cellulose

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Help: I'm working with cardiac cell motility in a methyl cellulose matrix. The
problem is that methyl cellulose does not lend itself readily to typical
specimen protcols for electron microscopy. Does anyone have a fixative or a
cross-linking chemical that will stabilize methyl cellulose at room temperatures
or cooler?

--

Darryl Krueger
Electron Optical Facility
University of Minnesota
Dept. Plant Pathology
495 Borlaug Hall
St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX
giba-at-puccini.crl.umn.edu






From: db0d-at-CS1.CC.Lehigh.EDU (DONNA BELCHER)
Date: Tue, 4 Oct 1994 12:35:27 -0400
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



subscribe




From: Nestor J. Zaluzec - Argonne Nat. Lab. :      ZALUZEC-at-ANLEMC.MSD.ANL.GOV
Date: Sat, 23 Jul 1994 20:58:16 -0500 (CDT)
Subject: Warning: ANLEMC is dying

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe microscopy elaine mohrbach








QQQQQ

---------- Forwarded message ----------


Fellow Microscopy Subscribers:

ANLEMC is dying..... I've managed to get it back up
and running, however, expect that sometime in the next
few days that this mail server will have a new home.

Please keep an eye on your mail for the announcement
of a new host address. I will have to permanently
move the microscopy mailing list so that the
problems of the last few days do not repeat themselves.

Sorry for the headaches many of you have had in the
last week, it wasn't fun for me either.

Nestor
-------------


Nestor J. Zaluzec
Argonne National Lab.
Materials Science Division






From: Greg Erdos ICBR EM Core Lab Univers :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Tue, 04 Oct 1994 15:10:56 -0500 (EST)
Subject: Safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


With the few brain cells I have left I have a dim memory of an extended
article on lab safety that may have appeared in either the EMSA Bulletin or in
the SEEMS
Newsletter before either changed their names. If anyone can point me in the
direction of this article or a similar one, I would be grateful.
As a corallary we would like to have some information on the hazards of
elemental Osmium as opposed to OsO4 (Safety dept wants to condemn our
refrigerator with the blackened interior. They intend to seal it up and have
it buried in a hazardous waste dump in Atlanta)
**********************************************************
* Greg Erdos ** *
* Director, ICBR EMCL ** Phone 904-392-1295 *
* 218 Carr Hall ** FAX 904-392-8598 *
* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
* Gainesville, FL 32611 ** *
**********************************************************




From: tivol-at-tethys.ph.albany.edu
Date: Tue, 04 Oct 1994 17:03:43 EDT
Subject: Safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Greg,
The Handbook of Chemistry and Physics says that bulk osmium is unaffec-
ted by air and is bluish-white; whereas the powdered metal reacts with air to
produce OsO4--the starting material in your case. I think you can argue that
the refrigerator is much safer than the contents. Only the oxide is toxic, and
it can be recognized by its smell. Good luck.
Yours,
Bill Tivol




From: Laurie Frederick :      frederick_laurie-at-vanlab.paprican.ca
Date: Tue, 4 Oct 1994 14:15:28 PDT
Subject: Re: Safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941004141528.416-at-vanlab.paprican.ca}
To: Greg Erdos ICBR EM Core Lab Univers {GWERDOS-at-gnv.ifas.ufl.edu}

There is a recent publication which might be of help to you entitled:

"Electron Microscopy Safety Handbook"
Vernon Barber and Joseph A. Mascorro, Editors
1994

published by San Franscisco Press Inc.
Box 426800,
CA. 94142-6800

ISBN 0-911302-72-7

I have just glanced through it but it had sections on Biological and
Materials preparation and X-ray safety as well.

Good Luck,
Laurie

} Date: Tue, 04 Oct 1994 15:10:56 -0500 (EST)
} From: Greg Erdos ICBR EM Core Lab Univers
{GWERDOS-at-gnv.ifas.ufl.edu}

} With the few brain cells I have left I have a dim memory of an
extended
} article on lab safety that may have appeared in either the EMSA
Bulletin or in
} the SEEMS
} Newsletter before either changed their names. If anyone can point
me in the
} direction of this article or a similar one, I would be grateful.
} As a corallary we would like to have some information on the
hazards of
} elemental Osmium as opposed to OsO4 (Safety dept wants to
condemn our
} refrigerator with the blackened interior. They intend to seal it up and
have
} it buried in a hazardous waste dump in Atlanta)

______________________________________________________________________
Laurie Frederick, A.SC.T. PAPRICAN
Corrosion Control Group 3800 Wesbrook Mall
The Pulp and Paper Research Vancouver, B.C.
Institute of Canada Canada V6S 2L9

Email: frederick_laurie-at-vanlab.paprican.ca
Tel: 604-222-3200 Fax: 604-222-3207




From: _ _ _ _ _ MARK W. LUND _ _ _ _ _ :      LUNDM-at-PHYSC1.BYU.EDU
Date: Tue, 4 Oct 1994 15:32 MDT
Subject: Osmium Safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Greg,
Osmium compounds are bad, but osmium is probably the most inert (or maybe
ruthenium is more inert) metal there is. The reason that osmium compounds
are bad is that they want to dump their osmium somewhere, in exchange for
whatever you offer. Don't let your bean counters take your fridge, just
because it is coated with precious osmium! And for heaven's sake don't
let them know that gold compounds are toxic. They will want to bury your
wedding ring with the fridge!

with tongue in cheek
Mark Lund




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Tue, 4 Oct 1994 18:17:55 -0400 (EDT)
Subject: Re: Long messages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am late to this string, because I was out of town for some time. I
agree that long messages cause problems for all, however, in the interest
of widest dissemination of information I would suggest also posting a
short message to the server indicating that additional information is
available directly. Just my HO.

Best-

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************


On Fri, 30 Sep 1994, L. D. Marks wrote:

} I have just removed a message of over 500 lines concerning
} Yeast Fixation. Can I make an appeal to people not to send so much
} detail straight to the server -- this type of information should
} go over more private channels directly between interested parties.
} Yeast fixation may be an important topic, but not everyone is
} interested is all the details.
}
} Thanks
}
}




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Wed, 5 Oct 1994 12:34:54 +1300
Subject: Drying plant material - Help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Could anyone help regarding processing of plant material. We have been
prcessing plants for SEM by conventional methods with 2.5% glutaraldehyde
in 0.1 M cacodylate buffer, osmication, dehydration and critical point
drying. We noticed that the plant material, whether leaf, grass or lichen
have a tendancy to explode into fragments during decompression at the end
of critical point drying. We thought maybe air was still trapped within the
palnt tissues and caused the samples to pop. We fix the tissue under vacuum
to remove air but still experiance the same problem.

Mark Gould
University of Otago
Dunedin
New Zealand

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: tony-at-emu.su.oz.au (Tony Romeo)
Date: Wed, 5 Oct 1994 10:45:57 +0100
Subject: Plant Material in the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410050040.AA01027-at-ELECTRON.emu.su.OZ.AU}
X-Sender: tony-at-electron.emu.su.oz.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

G'day Richard,
With regard to your 'exploding' plant material, I would
suggest that you slow down the venting procedure at the end of the CPD run.
Your specimens could be displaying a botanical version of the 'bends'!!?

As a possible alternative, if this is not the case, we've been having great
success recently with a 'low-tech' cold stage procedure. Basically it
involves sticking (with dag) a piece of fresh plant specimen onto a 2cmx1cm
brass block then plunging the whole thing into liquid nitrogen for about 45
- 60 secs ( this time is important - you don't want the specimen to get too
cold). The cold specimen is then quickly transferred to the SEM and you've
got viewing times of about 45 - 75mins. I've found that the more hydrated a
specimen is, the better it seems to work, i.e. the less charging it
displays, possibly because the water is acting as a conductor. This
technique is ideal if you're looking at external features and we've even
managed to remove cold specimens, fracture them, and then return them to
the microscope without the need for transfer devices etc. I've been knocked
out with the ease and the results of the technique - try it!

Some reading material along these lines can be found in

Microscopy Research and Technique 1994 vol 28 : 67 - 74

Journal of Microscopy 1993 vol. 172 : 63 - 69

Scanning 1993 vol. 15 : 171 - 173

Cheers, Tony Romeo

Tony Romeo Internet: tony-at-emu.su.oz.au
Electron Microscope Unit Telephone: 692 2351
Madsen Building F09
The University of Sydney Facsimlie: 552 1967
Sydney, 2006
Australia







From: IAN HALLETT :      ihallett-at-marccri.marc.cri.nz
Date: Wed, 5 Oct 1994 15:46:26 GMT+1200
Subject: Re-Drying plant material

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Richard and Mark

I would agree with Tony Romeo that your main problem may be speed of
venting. Our CPD system recommends not faster than 1-200psi per
minute and we err on the side of caution.

I note you use an osmium post fix. We normally just glutaraldehyde
fix before dehydration for plant material that we have to Critical
Point Dry although low temperature observation is much the best for
much of our work.

Ian
Ian Hallett
HortResearch
Mt Albert Research Centre
Private Bag 92 169
Auckland, New Zealand
Fax 64-9-815 4201
Telephone 64-9-849 3660




From: Luc Analbers :      L.J.S.Analbers-at-med.ruu.nl
Date: Wed, 05 Oct 1994 09:57:36 +0100
Subject: Cell-cultures and Immuno-EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi,

I have some serious problems dealing with immuno-EM and embedding.
We would like to embed endothelial cells (cultured) and afterwards do an
incubation (immunogold).
First, we've tried Unicryl to embed the cells cultured in polystyrene
dishes (Falcon). This couses severe problems because Unicryl melts the
polystyrene dishes.
I think that Lowicryl will raise the same problems.
So we have tried to embed the cells in Araldite. The embedding was succesfull,
but we didn't succeed in getting a nice immunolabeling.

IS ANYONE EXPERIENCED IN EMBEDDING CULTURED CELLS FOR IMMUNO-EM?
We would like to keep the cell layer intact (and not scraping the cells for
suspensions).

Thanks.


Luc.

***************************************************************************
* Luc Analbers * Analbers-at-med.ruu.nl *
***************************************************************************
* Utrecht University * LLL *
* Medical Faculty * LLL *
* Dept. Medical Physiology & * LLL A *
* Sportsmedicine * LLL AA AA *
* PO-box 80043 * LLL AA AA *
* Zip: 3508 TA * LLLLLAAALLLAAALLL *
* Utrecht * LLLLLAAALLLAAALLLL *
* Tel: 030 - 538911 * AAA AAA *
* Fax: 030 - 539036 * AAA AAA *
***************************************************************************






From: W.L. Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Wed, 5 Oct 1994 08:40:45 EST
Subject: Cell culture & IEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To Luc Analbers,

We routinely process intact monolayers for IEM. I early abandoned
polystyrene dishes for the same reasons you mentioned. I recommend to all
our users to culture cells in Leighton tubes...these are flat-bottomed
screw-cap tubes that lie horizontal. The cells grow on an easily-detached
plastic coverslip that survives any chemical processing, even propylene
oxide. Before embedding, the coverslip can be cut up, flat embedded, and
sectioned along with the attached monolayer. It actually sections better
than the LR White resin that we normally use.
Leighton tubes are available from most of the suppliers of plasticware
for cell culture. Hope this helps.

-=W.L. Steffens=-
College of Veterinary Medicine
University of Georgia




From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Wed, 5 Oct 1994 08:48:50 -0500
Subject: Re: Cell-cultures and Immuno-EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We routinely embed monolayers growing on glass coverslips in Falcon or
Costar 12 well trays using Epon, LR Gold, and JB-4. In the past, I have
used Lowicryl K4M and HM20 but I am unsure what type of trays we were using
back then. We never have any problem with the trays dissolving. We use
only ethanol for dehydration; no acetone or propylene oxide. We have had
some problems with LR White doing weird things which I guess (without a
whole lot of confidence) is due to differential contraction of the plastic
as it polymerizes. Once polymerized, we cut out the bottom of the well
using a Dremel moto-tool, cross-hatch the thin layer of plastic above the
glass coverslip and slowly immerse in liquid nitrogen. the plastic
contracts at a different rate than the glass coverslip and the cells pop
off with the plastic in nice little flat squares. we used to re-embed the
squares in standard molds but now simply cut them immediately with
virtually no trimming needed. it works great. contact me directly if you
have any questions. good luck.


Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: MILLERS-at-NCCCOT.AGR.CA
Date: 05 Oct 1994 11:55:12 -0500 (EST)
Subject: source for Kodak dyes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello microscopists!!
I would very much like to order some dyes/stains from Eastman Kodak,
but can't seem to find a valid address/phone/fax #. Can anyone out there
help me??
Thanks in advance,

Shea Miller
Agriculture and Agri-Food Canada
Ctr. for Food and Animal Research
Rm. 2002, K.W. Neatby Bldg.
Central Experimental Farm
Ottawa, Ontario
Canada K1A 0C6
millers-at-ncccot.agr.ca




From: MELSEN :      MELSEN-at-microbio.emory.edu
Date: 5 Oct 1994 13:48:37 EST
Subject: Re: source for Kodak dyes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Shea,

Try:
Eastman Kodak Co.
Eastman Fine CHemicals
343 State Street
B 701
Rochester, N.Y. 14652-3512

800 225 5352
FAX 716 722 3179

Regards, Skip
melsen-at-MICROBIO.emory.edu




From: morilak-at-thorin.uthscsa.edu (David Morilak)
Date: Wed, 5 Oct 1994 13:26:31 -0600
Subject: Re: micromanipulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Norman:

You might try Narishige - they are one of the finest sources of
micromanipulators, at least for electrophysiologic applications. I don't
have a number at hand, but they should be easy to track down - they will be
listed in the Science Guide to Scientific Products, and also in the vendor
directory in the Program of the Society for Neuroscience meeting (also I
imagine FASEB, IBRO and many others). Good luck!


David Morilak
Dept Pharmacology
Univ Texas Health Science Center
San Antonio

ph: 210-567-4174
FAX: 210-567-4303
E-mail: morilak-at-uthscsa.edu






From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Wed, 5 Oct 1994 15:59:34 -0400 (EDT)
Subject: RE: source for Kodak dyes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Wed, 5 Oct 1994 12:55:12 -0400, MILLERS-at-NCCCOT.AGR.CA writes:

} Hello microscopists!!
} I would very much like to order some dyes/stains from Eastman Kodak,
} but can't seem to find a valid address/phone/fax #. Can anyone out there
} help me??
} Thanks in advance,
}
} Shea Miller
} Agriculture and Agri-Food Canada
} Ctr. for Food and Animal Research
} Rm. 2002, K.W. Neatby Bldg.
} Central Experimental Farm
} Ottawa, Ontario
} Canada K1A 0C6
} millers-at-ncccot.agr.ca
}
=======================

Kodak Laboraory Chemicals suppliers in Ontario, Canada:

Anachemia Science Anachemia Science
2708 Southview Drive Unit A, 3120 Pepper Mill Court
Box Site 31, Box 10 Misissauga, Ontario L5L 4X4
Sudbury, Ontario P3E 4M9 Tel: 416-82804409
Tel: 705-522-5501

Fisher Scientific Ltd.
1200 Denison Street
Unionville, Ontario L3R 8G6
Tel: 416-479-8700

*****************************************************

M.V. Parthasarathy
Section of Plant Biology, 228 Plant Science Building
Cornell University, Ithaca, NY 14853. USA
Tel: 607-255-1734; Fax: 607-255-5407
E-Mail: mvp2-at-cornell.edu

******************************************************




From: Elaine Humphrey :      ech-at-unixg.ubc.ca
Date: Wed, 5 Oct 1994 14:51:49 -0700 (PDT)
Subject: Re: Cell-cultures and Immuno-EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Have you tried growing your cells on Aclar plastic film instead of
polystyrene dishes? Embedding in Lowicryl HM20 works well.
Elaine

On Wed, 5 Oct 1994, Luc Analbers wrote:

}
} Hi,
}
} I have some serious problems dealing with immuno-EM and embedding.
} We would like to embed endothelial cells (cultured) and afterwards do an
} incubation (immunogold).
} First, we've tried Unicryl to embed the cells cultured in polystyrene
} dishes (Falcon). This couses severe problems because Unicryl melts the
} polystyrene dishes.
} I think that Lowicryl will raise the same problems.
} So we have tried to embed the cells in Araldite. The embedding was succesfull,
} but we didn't succeed in getting a nice immunolabeling.
}
} IS ANYONE EXPERIENCED IN EMBEDDING CULTURED CELLS FOR IMMUNO-EM?
} We would like to keep the cell layer intact (and not scraping the cells for
} suspensions).
}
} Thanks.
}
}
} Luc.
}
} ***************************************************************************
} * Luc Analbers * Analbers-at-med.ruu.nl *
} ***************************************************************************
} * Utrecht University * LLL *
} * Medical Faculty * LLL *
} * Dept. Medical Physiology & * LLL A *
} * Sportsmedicine * LLL AA AA *
} * PO-box 80043 * LLL AA AA *
} * Zip: 3508 TA * LLLLLAAALLLAAALLL *
} * Utrecht * LLLLLAAALLLAAALLLL *
} * Tel: 030 - 538911 * AAA AAA *
} * Fax: 030 - 539036 * AAA AAA *
} ***************************************************************************
}
}
}




From: {Luc Analbers :      L.J.S.Analbers-at-med.ruu.nl }:unix:niaid
Date: 10-5-1994 9:57am
Subject: Cell-cultures and Immuno-EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: cfg:rml:niaid,
sfh:rml:niaid,
dwd:rml:niaid
Subj: Cell-cultures and Immuno-EM
Also-to: microscopy-at-aaem.amc.anl.gov

--------------------------------------------------------------------


Hi,

I have some serious problems dealing with immuno-EM and embedding.
We would like to embed endothelial cells (cultured) and afterwards do an
incubation (immunogold).
First, we've tried Unicryl to embed the cells cultured in polystyrene
dishes (Falcon). This couses severe problems because Unicryl melts the
polystyrene dishes.
I think that Lowicryl will raise the same problems.
So we have tried to embed the cells in Araldite. The embedding was succesfull,
but we didn't succeed in getting a nice immunolabeling.

IS ANYONE EXPERIENCED IN EMBEDDING CULTURED CELLS FOR IMMUNO-EM?
We would like to keep the cell layer intact (and not scraping the cells for
suspensions).

Thanks.


Luc.

***************************************************************************
* Luc Analbers * Analbers-at-med.ruu.nl *
***************************************************************************
* Utrecht University * LLL *
* Medical Faculty * LLL *
* Dept. Medical Physiology & * LLL A *
* Sportsmedicine * LLL AA AA *
* PO-box 80043 * LLL AA AA *
* Zip: 3508 TA * LLLLLAAALLLAAALLL *
* Utrecht * LLLLLAAALLLAAALLLL *
* Tel: 030 - 538911 * AAA AAA *
* Fax: 030 - 539036 * AAA AAA *
***************************************************************************







From: Alex :      ALEX-at-bunyip.ph.rmit.oz.au
Date: 6 Oct 94 13:53:31 EST-10
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941006135331.384-at-bunyip.ph.rmit.edu.au}
To: microscopy-at-aaem.amc.anl.gov

subscribe microscopy Alex Titkov




From: marilyn-at-cemmsa.adelaide.edu.au (Marilyn Henderson)
Date: Thu, 6 Oct 1994 13:35:17 +0900
Subject: Cell-culture and Immuno-EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199410060405.EAA15770-at-traminer.cemmsa.adelaide.edu.au}
X-Sender: marilyn-at-traminer.cemmsa.adelaide.edu.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"



I have had similar problems in the past with cell cultures brought to me on
plastic microwell dishes or cell-culture flasks. Dehydration with ethanol
is not a problem but EVERY resin I've tried- Spurrs, Taab, LR White, HM20 -
has melted the plastic. I usually advise those growing the cells to use
Thermanox coverslips or polycarbonate filters on which to grow the cells as
these can be processed and cut with the cells; thus solving many problems.
However, some researchers still find it necessary to use platic (or they
alrady have the cells grown before I'm consulted). Thus, I have discovered
that by exposing the cells and dish to HM20 for 1 hour at room temperature
the dish will melt sufficiently to allow the cells to float off into the
resin.(You need an intact monalyer though). I then lift out sections of the
cell monolayer and put them in fresh HM20 and polymerize at room
temperature under UV light.
HN20 will give excellent immunolabelling results with immunogold.

Marilyn Henderson
Biological Electron Microscopist
CEMMSA
The University of Adelaide
South Australia






From: Loling Song :      loling-at-ruly46.LeidenUniv.nl
Date: Thu, 6 Oct 1994 09:30:42 +0000 (WET)
Subject: Re: source for Kodak dyes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On 5 Oct 1994 MILLERS-at-NCCCOT.AGR.CA wrote:

} Hello microscopists!!
} I would very much like to order some dyes/stains from Eastman Kodak,
} but can't seem to find a valid address/phone/fax #. Can anyone out there
} help me??
} Thanks in advance,
}
} Shea Miller
} Agriculture and Agri-Food Canada
} Ctr. for Food and Animal Research
} Rm. 2002, K.W. Neatby Bldg.
} Central Experimental Farm
} Ottawa, Ontario
} Canada K1A 0C6
} millers-at-ncccot.agr.ca
}

The Kodak distributor near you is:

Fisher Scientific, Ltd.,
1200 Denison Street,
Unionville, Ontario L3R 8R6
tel: 416-479-8700
fax: 416-479-9749

or call toll-free:
800-225-5352 (Kodak)

I hope that this information is still up-to-date.

Success!

Loling Song

===============================================================================
Loling Song Office tel: +31+71-276198
Department of Cytochemistry and Cytometry, +31+71-276200
Leiden University, fax: +31+71-276180
Wassenaarseweg 72,
2333 AL Leiden,
The Netherlands email: song-at-RULLF2.LeidenUniv.NL
===============================================================================





From: Felix De La IgIesia 7349 :      delaigf-at-aa.wl.com
Date: Thu, 06 Oct 1994 07:20:33 -0400 (EDT)
Subject: Unsuscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Unsuscribe delaigf%am%petvax-at-mr.research.aa.wl.com






From: Mike Witcomb :      MIKEW-at-gecko.biol.wits.ac.za
Date: Thu, 6 Oct 1994 13:28:55 GMT+2
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe Microscopy mikew-at-gecko.biol.wits.ac.za




From: emlab-at-ucsco.ucsc.edu (Jon Krupp)
Date: Thu, 6 Oct 1994 08:36:54 -0700
Subject: Address for BEEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can anyone bring me up to date on an address or phone number for BEEM?. I
tried an old one (212-597-3670) but it did not work. I want to find out
if their developing tank system is still available.

Thanks

Jonathan Krupp
Microscopy and Imaging Lab
University of California
Santa Cruz, CA 95060
emlab-at-ucsco.ucsc.edu
(408) 459-2477






From: _ _ _ _ _ MARK W. LUND _ _ _ _ _ :      LUNDM-at-PHYSC1.BYU.EDU
Date: Thu, 6 Oct 1994 20:30 MDT
Subject: tv coat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Last year I bought a colloidal carbon paint called TV coat from
Ted Pella. This year they no longer supply it. I need some, does
anyone know where to get it?
regards
Mark W. Lund
MOXTEK, Inc.
Orem UT




From: Stephan Coetzee :      STEPHAN-at-gecko.biol.wits.ac.za
Date: Fri, 7 Oct 1994 09:02:10 GMT+2
Subject: SUBSCRIBE MICROSCOPY

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


SUBSCIBE MICROSCOPY (STEPHAN-at-GECKO.BIOL.WITS.AC.ZA)
Stephan H Coetee
Electron Microscope Unit
Private Bag 3
Wits
2050

Tell: (011) 716 2419
Fax : (011) 339 3407




From: loewe-at-uni-bonn.de (Andreas Loewe)
Date: Fri, 7 Oct 1994 09:02:10 GMT+2
Subject: address of "The American Ceramic Society"?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: UNC900-at-ibm.rhrz.uni-bonn.de
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: Microscopy-at-aaem.amc.anl.gov

Please help,
I need to know how to access "The American Ceramic Society". I applied for
membership allready some month ago but the only reaction I got was a
booking on my mastercard account.
Is there any way to access them by email? Their faxmachine seems to be
somewhere in the basement.

Any help highly appreciated!!!

Andreas Loewe

_________________________________________________________________
Andreas Loewe **** netadmin **** Phone: +49-228-550325
Anorganische Materialforschung +49-228-550204
Institut fuer Anorganische Chemie Fax: +49-228-678413
der Universitaet Bonn email: loewe-at-uni-bonn.de
Roemerstrasse 164 ____ _
D-53117 Bonn //_// //__
Germany // // /___/
_________________________________________________________________






From: paul-at-etlcom3.etl.go.jp (Paul Fons)
Date: Fri, 7 Oct 1994 20:22:58 +0900
Subject: Capping Layer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410071122.AA21822-at-etlpom.etl.go.jp}

We have some chalcopyrite material (CuInSe2) I am looking at here using
cross sectional TEM and I would really like to focus on the near surface
(several monolayers). I suspect that the top few layers would never
survive the cross sectional sample preparation process so I was wondering
about the idea of putting an ex-situ cap on it (as immediately after growth
as possible) of some non-reactive material (SiN would be one example). I
realize this might generate some stress in the material, but it is all I
can think of at the moment. We can't put anything on it in situ that won't
have the possibility of reaction with it (Se4 doesn't stick significantly
at room temperature). Any thoughts on this or suggestions as to what might
be a good capping layer.





From: W.L. Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Fri, 7 Oct 1994 08:51:17 EST
Subject: TV Tube Koat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mark,
TV Tube Koat is generally available from electronics supply houses.
It is used to maintain conductivity between vacuum tube plugs and the
socket. The manufacturer is:

G.C. Electronics
Division of Hydrometals, Inc.
Rockford, IL 61101

Let me know if you are successful...I need some too.

-=W.L. Steffens=-
College of Veterinary Medicine
University of Georgia




From: POSHEL-at-wpo.it.luc.edu
Date: Fri, 07 Oct 1994 08:37:54 -0500
Subject: Drying plant material - Help -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {se950944.017-at-wpo.it.luc.edu}
X-Mailer: WordPerfect Office 4.0
richard.easingwood-at-stonebow.otago.ac.nz

Have you tried chemical drying with Pel-Dri or hexamethyldisilizane?
After 100% EtOH, go through a 1:2, 1:2, 2:1 series EtOH:chemical
then 3 changes in HMDS or P-D
then dry: sublimate Pel-Dri in a vacuum dessicator under mild! vacuum
(just enough to remove the vapors) below 22 C
or dry HMDS in a fume hood overnight or in a fume hood for 2 hr at 60 C
neither may be as good as observing frozen-hydrated, but neither do
they require a cold-stage.
Phil "looking for a job" Oshel
poshel-at-luc.edu






From: Sally Shrom :      sally-at-retina.anatomy.upenn.edu
Date: Fri, 7 Oct 1994 10:20:43 -0400 (EDT)
Subject: acrolein

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Posted-Date: Fri, 7 Oct 1994 10:19:59 -0400

Has anyone ever used acrolein? I must use it in a fixative,and I am
afraid of it. What special precautions must I take. I am going to use it
in a fume hood,and I will wear my standard protective gloves. Will that
be good enough or do I need to dress up like an astronaut?




From: {pinglu-at-alumina.rutgers.edu}:ddn:wpafb
Date: 10-6-94 12:19pm
Subject: Re: program to calculate grain misorientation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410071446.AA09943-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: program to calculate grain misorientation
Orig-Author: {pinglu-at-alumina.rutgers.edu (Ping Lu)}:ddn:wpafb
-----------------------------------------------------------
Dear microscopists:

I am looking for a program to calculate grain orientation and misorientation
at grain boundary, based on electron beam kikuchi pattern. This program is
probably available there, I just do not know where to get it. Everyone knows,
please let me know.

Many thanks !

P. Lu
pinglu-at-alumina.rutgers.edu
Rutgers, N





From: Joyce Craig :      bafpjec-at-uxa.ecn.bgu.edu
Date: Sat, 8 Oct 1994 15:11:34 -0500 (CDT)
Subject: free cryoultramicrotomy workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Midwest Society of Electron Microscopists (MSEM), RMC of Tucson,
Arizona, and Integrated Microscytems are sponsoring a workshop
"Cryotechniques in Electron Microscopy" Friday, October 21, 1994 at
Chicago State University.

Program:
9:30AM Registration
10:00 AM Gregory M. Becker will speak on glass knife making for
cryo-ultramicrotomy and room temperature Ultramicrotomy and
cryo-Preparation Techniques including Metal-Block Freezing and Propane
Jet-Freezing.
11:00 AM Michelle Wilhite will speak on Practical Aspects of
Cryomicrotomy and Immunolabeling
12:00 Noon Luncheon ($7.00)
1:00 to 5:00 PM
Hands-on Workshop demonstrating cryosectioning of biological materials,
harvesting sections, and demonstration of tools
Demonstration of tools and techniques of immunolabeling and postembedding
Demonstration of metal-block freequing and other techniques
Facility Tour

Do you have any problems now? Bring them in for individual help.

The workshop is free to members of MSEM. Ask for an application.
Membership is only $10 per year, $5 for students

To reserve your space call now:
Steve Miller
FAX 708-696-2541
E-Mail 73150.2217-at-compuserve.com
Phone 1-800-388-8801




From: W.L. Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Sun, 9 Oct 1994 17:08:05 EST
Subject: Acrolein

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sally,
I have used acrolein for many years. It is an excellent fixative when
rapid penetration or stabilization lipids is required. You should be
aware of a few things about it. 1. It is flammable...be sure your hood
can handle this and there are no flames or sparks about. 2. It is very
volatile...thats what makes it useful as a vapor fixative. 3. It is very
irritating to mucous membranes...treat it as you would osmium. In the
concentrations that you are likely to use it in working solutions, its
really no worse than glutaraldehyde. Using good safety procedures, we
have never had an problems with it. It has a very distinctive odor (like
burning, rancid, fat), so you always know if you're inhaling it.

-=W.L. Steffens=-
College of Veterinary Medicine
University of Georgia




From: ZWANG-at-ENH.NIST.GOV
Date: Sun, 09 Oct 1994 21:32:20 -0400 (EDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe zwang-at-enh.nist.gov




From: tony-at-emu.su.oz.au (Tony Romeo)
Date: Mon, 10 Oct 1994 17:30:37 +0100
Subject: Low Tech Cryo-SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


G'day Gib,
Thanks for your comments. I would like to make it clear that the technique
I had outlined in my earlier note is not mine - it seems to be a procedure
which has been around for a while (in various levels of sophistication) but
it's value probably hasn't been appreciated as much as it should be. That's
certainly true in my case!

As for the kV's that we use, that's been very much a trial and error
procedure. On our JEOL 35C we are basically restricted to 10kv or above.
This has been OK for the plant material (15kv) and some of the insect stuff
(15 - 20kv). I have had to do some higher resolution work on latex
particles and for this we went to 20 - 25kv (for images at 40 - 50k). On
our Philips 505 we can obviously work at much lower kv's but it's also in
use most of the time so I haven't had much of an opportunity to 'play' on
it. For the things that I have done on it (latex particles) I've needed
higher resolution and so have been working up around the
20kv range.

Hope all that is of some use!

Cheers, Tony Romeo


Tony Romeo Internet: tony-at-emu.su.oz.au
Electron Microscope Unit Telephone: 692 2351
Madsen Building F09
The University of Sydney Facsimlie: 552 1967
Sydney, 2006
Australia







From: Sally Shrom
Date: Friday, October 07, 1994 10:20AM
Subject: acrolein

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Has anyone ever used acrolein? I must use it in a fixative,and I am
afraid of it. What special precautions must I take. I am going to use it
in a fume hood,and I will wear my standard protective gloves. Will that
be good enough or do I need to dress up like an astronaut?




From: Susan Smith - National Steel Corporation :      smiths-at-mlc.lib.mi.us
Date: Mon, 10 Oct 1994 11:32:39 -0400 (EDT)
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe smiths-at-mlc.lib.mi.us





From: Felix De La IgIesia 7349 :      delaigf-at-aa.wl.com
Date: Mon, 10 Oct 1994 16:53:36 -0400 (EDT)
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe delaigf-at-aa.wl.com






From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Mon, 10 Oct 1994 17:32:46 -0500
Subject: Re: image analysis for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
I know of two very nice and very inexpensive pieces of
software for doing image analysis on PC. Only trouble is, they each require
the purchase of a frame grabber card. The frame grabber card which they
need costs about 2,000 US dollars. The frame grabber card is the pcvision
plus card, and the software is: Morpho.sys and MTV. The software is in the
300 to 500 dollar range, and does stuff like angles, areas, lengths,
contours. ehgo.

Addresses/further info available on request.

AE
Tobias Baskin

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
___ ____ ^ ____ _____ Tobias I. Baskin
/ \ / / \ / \ / University of Missouri
/ | / / \ / / Biological Sciences
/___ / /__ /_____\ / /__ 109 Tucker Hall
/ / / \ ( / Columbia, MO 65211 USA
/ / / \ \ / voice: 314-882-0173
/ /____ / \ \____/ /_____ fax: 314-882-0123






From: IAN HALLETT :      ihallett-at-marccri.marc.cri.nz
Date: Tue, 11 Oct 1994 13:34:30 GMT+1200
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} From: forbes-at-cip.org.ec (Greg Forbes)
} Date sent: Mon, 10 Oct 94 16:28:20 EST
} To: microscopy-at-aaem.amc.anl.gov
} Subject: Re: Image Analysis Software for PC

}
} } Are there any image analysis software packages available for the PC that are
} } worth having? I need to do some fairly basic image analysis (such as area,
} } circumference etc.) on a PC system.
} }
} } Thanks
} }
} } Joe Michael
} } Sandia National Labs
} } jrmicha-at-sandia.gov
}
} Joe,
}
} I understand that there are two which are quite good and quite
} expensive: Optimas and Visilog. If you have a large budget and want
} addresses let me know. I would also like to know if there are any equal
} to these which are more economical.
}
} Please let me know if you get any answers which are not posted.
}
} Good luck,
}
} Greg
}
} --
} Greg Forbes Telfs. off: +593-2-690990
} International Potato Center fax: +593-2-562286
} P.O. 17-16-129 CEQ, Quito, Ecuador home: +593-2-330971
} Internet: forbes-at-cip.org.ec;
}
}

I have used both Optimas and Visilog and at the time I was assessing
Optimas was much better at basic measurements whilst Visilog was
a much more powerful at image processing. We now use Optimas
routinely. In New Zealand it cost around one third the price of
Visilog (say $US3000-$US4000). We have had a user working with JAVA
from Jandel Scientific who was very happy with it for basic measures,
this cost less than either Optimas or Visilog. I understand they
also produce a Windows image analysis package.

Unfortunately I have not found a satisfactory public domain imaging
package for the PC that was any match for NHI Image for the MAC.

Ian





Ian Hallett
HortResearch
Mt Albert Research Centre
Private Bag 92 169
Auckland, New Zealand
Fax 64-9-815 4201
Telephone 64-9-849 3660




From: Doug Arrell :      ARRELL-at-jrc.nl
Date: Tue, 11 Oct 1994 08:28:41 GMT+0200
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941011082841.544-at-FS-IAM-1.JRC.NL}

} Are there any image analysis software packages available for the PC that are
} worth having? I need to do some fairly basic image analysis (such as area,
} circumference etc.) on a PC system.
}
I know this sounds negative, but I wouldn't bother. My recomendation
would be to by a Macintosh and download NIH-Image from
ZIPPY.NIMH.NIH.GOV
as this will cost you roughly the same as a 'cheap' commercial
package on a PC and in my (admittedly limited) experience is
superior. This software is very flexible and gives you virtually
direct e-mail access to the author via a user group if you have any
problems. The software is also very fast (for a non-hardware analysis
solution). For example I tested Image Pro Plus on a 33Mhz 486 PC
doing a 'sharpen' filter and this woked at roughly 40000 pixels per
second, whereas NIH Image on a PowerPC 8100 (80Mhz) managed about
710000 pix/sec.
Incidentally Image Pro Plus under windows has some bugs that I found
within 1 hr of starting to try it out, and I managed to crash my
system 3 times in 2 hours so I wouldn't recommend it.

Hope this helps ( :-) )

Doug Arrell

+------------------------------------+
| Dr Douglas Arrell |
| Mechanical Performance and Joining |
| Institute for Advanced Materials |
| 1755 ZG Petten |
| Netherlands |
| {ARRELL-at-JRC.NL} |
| Tel. (+31) 2246 5287 |
| Fax (+31) 2246 1917 |
+------------------------------------+




From: Keith R. Hallam :      k.r.hallam-at-bristol.ac.uk
Date: Tue, 11 Oct 1994 10:59:49 +0100 (BST)
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410110959.AA08607-at-zeus.bris.ac.uk}
discussion list)

} } Are there any image analysis software packages available for the PC that are
} } circumference etc.) on a PC system.
} }
} } Thanks
} } Sandia National Labs
} Joe,
}
} I understand that there are two which are quite good and quite
} expensive: Optimas and Visilog. If you have a large budget and want
} addresses let me know. I would also like to know if there are any equal
} to these which are more economical.
}
} Greg
}
I don't know how many people this might apply to, but it was only yesterday
I found out that our University (bristol, England) had a site licence for
Visilog. I had been sort of looking for low cost image analysis software
(SigmaScan had been brought to my attention) but nothing comes lower in price
than something I could get for free from our Computer Centre (manuals will
cost 30pounds). The University had paid, maybe, 2000 pounds for the licence
but it can then be installed on any number of systems around the University.
The most important thing, though, is that the ownership of this wasn't
widely advertised so it *could* be worth spending a few minutes on the 'phone
seeing if such arrangements exist in your institution.

Keith
}
}
}
}
}






From: Alan Pooley :      pooley-at-ahab.rutgers.edu
Date: Tue, 11 Oct 1994 09:55:29 -0400
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lets distinguish between image processing and image measurement,
Optimas is good at changing and manipulating images
and pretty good at multiple measuring, but if accurate, controlable
measurements of images is what you want, let me suggest you look into
Morphosys, sold by Exeter sofware 516 689 7838 for $250.00. They sell
it for the U of Calif Berkeley regents, it was written by Chris Meacham
of the Herbarium and is a very nice program for measurement. I much prefer
its simplicity to Optimas or to older prepc systems I have used.
I have no connection other than pleased user.
Alan Pooley Rutgers Marine science




From: Alan Pooley :      pooley-at-ahab.rutgers.edu
Date: Tue, 11 Oct 1994 10:03:36 -0400
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


PS to my previous meassage, as some one reminded me, It is necessary to
have a PCVision Board (Cost $2000? ) and a tv camera to use Morphosys,
but you can use the same board for Optimas if you need to "up"grade later
and will need some sort of image capture hardware for any system. I have had
good luck with the PCVision Board, which comes with diagnostic
software (if people have trouble, as I did, I can give them a patch using
the ofg that will preset the image board for use with Morphosys
Alan Pooley Rutgers Marine sci




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Tue, 11 Oct 1994 9:40:10 -0500 (CDT)
Subject: FTP Site

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



NIH image can be obtained by anonymous FTP from a variety of sites:

Zippy.Nimh.Nih.gov
WWW.AMC.ANL.GOV

The first site is NIH, the second is the Microscopy FTP site which
includes the EMMPDL, MASLIB and selected Public Domain Shareware....

---Nestor---




From: loewe-at-uni-bonn.de (Andreas Loewe)
Date: Tue, 11 Oct 1994 9:40:10 -0500 (CDT)
Subject: need server address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: UNC900-at-ibm.rhrz.uni-bonn.de (Unverified)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: Microscopy-at-aaem.amc.anl.gov

recently I saw someone posting the address of a server named:
ZIPPY....

unfortunately I missed the complete name, please repost

Andreas Loewe






From: Jim Stanley :      jstanly-at-mse.ogi.edu
Date: Tue, 11 Oct 1994 09:22:05 -0700 (PDT)
Subject: Quantitative Oxygen

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410111536.AA12527-at-MIT.EDU}


Recently, I have had an industrial client ask me about methods for
obtaining accurate measurement of oxygen in a cobalt alloy in the 100 to
200 ppm range with some reasonable resolution...say microns. The sample
is a cross-section and the interest is in oxygen as a function of distance
from the surface. I've talked with some local talent with access to a SEM
equipped with WDS capability but it doesn't sound too promising. Any
suggestions??

Thanks in advance.


James T. Stanley, II
Oregon Graduate Institute
Portland, OR
jstanly-at-mse.ogi.edu






From: sbarlow-at-sunstroke.sdsu.edu (Steve Barlow)
Date: Tue, 11 Oct 1994 09:38:00 -0700
Subject: osmium as arthritis treatment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410111710.AA16711-at-MIT.EDU}

To those medical archivists among you, can anyone out there confirm this
anectodal story? I would particularly enjoy any references from texts or
articles.

I heard years ago from John Luft that osmium had been used as a
'treatment' for arthritis. Osmium was injected into the painful joint of
arthritis sufferers, whereupon it fixed the nerve cells. thereby treating
the symptoms. Becuase of its poor diffusion rate and and rapid reduction,
it was immobilized at the site of injection and did not travel elsewhere
in the body.

(Thank goodness we now have some better treatments.)

thanks in advance

steve



----------------------------------------------------------
Dr. Steven Barlow
EM Facility/Biology Dept.
San Diego State University
San Diego CA 92182-0057
phone: (619) 594-4523
fax: (619) 594-5676
email to sbarlow-at-sunstroke.sdsu.edu





From: Alan Pooley :      pooley-at-ahab.rutgers.edu
Date: Tue, 11 Oct 1994 12:31:49 -0400 (EDT)
Subject: image measurement program for pc

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reply to question about an image measurement program for the PC

I use Morphosys and like it very much, in preference to Optimas
at 10 times the price. However Morphosys does no do image
processing, ie filters etc nor does it trace/count/etc multiple
image objects without individual mouse clicks to direct it.
It does a wide variety of geometric steps as well as tracing well
contrasted objects amost instantly.
(If people have trouble getting good contrast to trace images ie can't get
the 'dark field image to follow the edge of the object, I can give
some hints about annual light, apertures to define edges and vanes to
control fine lighting details, using cardboard and a cheap circular
florescent light)
The program Morphosys costs $250.00 from Exeter Software 800 842 5892
It is property of U of Calif and written by Chris Meacham of the Herbarium
there. I have no connection other that as very satisfied user.
It does require the PCVision frame grabber board, at ??$2000.00 800 532 3500
and a tv camera and stand to hold it and the light etc.
All output of Morphosys is text and easily edited, rearranged, portions
deleted etc before geometric computations done and output put out.
alan pooley rutgers marine science





From: {jstanly-at-mse.ogi.edu}:ddn:wpafb
Date: 10-11-94 1:10pm
Subject: Re: Quantitative Oxygen

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: Quantitative Oxygen
Orig-Author: {Jim Stanley {jstanly-at-mse.ogi.edu} }:ddn:wpafb
-----------------------------------------------------------

Recently, I have had an industrial client ask me about methods for
obtaining accurate measurement of oxygen in a cobalt alloy in the 100 to
200 ppm range with some reasonable resolution...say microns. The sample
is a cross-section and the interest is in oxygen as a function of distance
from the surface. I've talked with some local talent with access to a SEM
equipped with WDS capability but it doesn't sound too promising. Any
suggestions??

Thanks in advance.


James T. Stanley, II
Oregon Graduate Institute
Portland, OR
jstanly-at-mse.ogi.edu






From: ard-at-atom.ansto.gov.au (Arthur Day)
Date: Wed, 12 Oct 1994 14:39:39 +1000
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {aac0176f00021004f06f-at-[137.157.15.210]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Are there any image analysis software packages available for the PC that are
} worth having? I need to do some fairly basic image analysis (such as area,
} circumference etc.) on a PC system.
}
} Thanks
}
} Joe Michael
} Sandia National Labs
} jrmicha-at-sandia.gov


There was some discussion about PC image analysis software a few weeks ago
on the NIH-Image mailing list. One of the replies mentioned the existence
of a free image analysis program for the PC called UCFIMAGE. The reply
stated that it was "developed by Weeks and Mylar at UCF ..... and you can
access it via anonymous ftp from IPG.ENGR.UCF.EDU in the /PUB/UCFIMAGE
directory".

I have not tried it myself because we use the Mac based NIH-Image program,
which serves our needs admirably.

Arthur Day, Electron Microscopy Group
Ansto Advanced Materials Program Phone: 61-2-717-3457
PMB 1, Menai (Sydney), NSW, 2234 Fax: 61-2-543-7179
Australia
Email: ard-at-atom.ansto.gov.au







From: Doug Arrell :      ARRELL-at-jrc.nl
Date: Wed, 12 Oct 1994 08:46:58 GMT+0200
Subject: NIH Image sites

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941012084658.704-at-FS-IAM-1.JRC.NL}

In response to the various questions about accessing NIH Image:

The most up to date versions are available on

ZIPPY.NIH.GOV (a shorter name than the one I gave yesterday)

via anonymous FTP.

To find the files look in the PUB/IMAGE directory.
In there you will find 4 versions of the software. Version 1.55 is
the most stable and is available in floating point or non-floating
point version along with the relevant manuals (word 4 mac format) and
even the souce code as the compressed files

NIH-IMAGE155_DOCS.HQX
NIH-IMAGE155_FPU.HQX
NIH-IMAGE155_NONFPU.HQX
NIH-IMAGE155_SOURCE.HQX

There is also a slightly improved version without documentation
NIH-IMAGE156BETA18

and a very fast but buggy beta for PowerPCs
NIH-IMAGE156PPC_BETA28

I hope this information is enough for you.

There is also a user group which is very helpful. To subscribe send a
message to LISTSERVER-at-UMN.SOILS.EDU with the line

SUBSCRIBE NIH-IMAGE your-at-address

Good luck

Doug Arrell

+------------------------------------+
| Dr Douglas Arrell |
| Mechanical Performance and Joining |
| Institute for Advanced Materials |
| 1755 ZG Petten |
| Netherlands |
| {ARRELL-at-JRC.NL} |
| Tel. (+31) 2246 5287 |
| Fax (+31) 2246 1917 |
+------------------------------------+




From: Petr Schauer :      petr-at-ISIBrno.Cz
Date: Wed, 12 Oct 1994 08:45:27 +0100
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} There was some discussion about PC image analysis software a few weeks ago
} on the NIH-Image mailing list. One of ...............
^^^^^^^^^^^^^^^^^^^^^^

Please, what is the listserver address of this NIH-Image list?

Petr
+-------------------------------------------+-------------------------+
| Dr. Petr Schauer | tel.: (+42 5) 41321246 |
| ACADEMY OF SCIENCES OF THE CZECH REPUBLIC | fax : (+42 5) 41211168 |
| INSTITUTE OF SCIENTIFIC INSTRUMENTS | E-mail: petr-at-isibrno.cz |
| Kralovopolska 147, CZ-612 64 Brno | |
| Czech Republic | |
+-------------------------------------------+-------------------------+




From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Wed, 12 Oct 1994 08:51:11 -0500
Subject: Re: osmium as arthritis treatment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} To those medical archivists among you, can anyone out there confirm this
} anectodal story? I would particularly enjoy any references from texts or
} articles.
}
} I heard years ago from John Luft that osmium had been used as a
} 'treatment' for arthritis. Osmium was injected into the painful joint of
} arthritis sufferers, whereupon it fixed the nerve cells. thereby treating
} the symptoms. Becuase of its poor diffusion rate and and rapid reduction,
} it was immobilized at the site of injection and did not travel elsewhere
} in the body.
}
} (Thank goodness we now have some better treatments.)
}
} thanks in advance
}
} steve


The reference for this work is Chem. & Eng. News 60:8 (April 5, 1982).

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: Doug Arrell :      ARRELL-at-jrc.nl
Date: Wed, 12 Oct 1994 16:13:07 GMT+0200
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941012161307.832-at-FS-IAM-1.JRC.NL}

Sorry, the listserver address for NIH-Image should have been

LISTSERVER-at-SOILS.UMN.EDU

not

LISTSERVER-at-UMN.SOILS.EDU

It was too early in the morning when I wrote the original message!

Doug Arrell

+------------------------------------+
| Dr Douglas Arrell |
| Mechanical Performance and Joining |
| Institute for Advanced Materials |
| 1755 ZG Petten |
| Netherlands |
| {ARRELL-at-JRC.NL} |
| Tel. (+31) 2246 5287 |
| Fax (+31) 2246 1917 |
+------------------------------------+




From: db0d-at-CS1.CC.Lehigh.EDU (DONNA BELCHER)
Date: Wed, 12 Oct 1994 11:13:49 -0400
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



unsubscribe db0d-at-cs1.cc.lehigh.edu




From: Bo Johansen :      BOJ-at-bot.ku.dk
Date: Wed, 12 Oct 1994 16:48:35 GMT+0100
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Doug Arrell wrote:

} I know this sounds negative, but I wouldn't bother. My recomendation
} would be to by a Macintosh and download NIH-Image from
} ZIPPY.NIMH.NIH.GOV
} as this will cost you roughly the same as a 'cheap' commercial
} package on a PC and in my (admittedly limited) experience is
} superior. This software is very flexible and gives you virtually
} direct e-mail access to the author via a user group if you have any
} problems. The software is also very fast (for a non-hardware analysis
} solution). For example I tested Image Pro Plus on a 33Mhz 486 PC
} doing a 'sharpen' filter and this woked at roughly 40000 pixels per
} second, whereas NIH Image on a PowerPC 8100 (80Mhz) managed about
} 710000 pix/sec.
} Incidentally Image Pro Plus under windows has some bugs that I found
} within 1 hr of starting to try it out, and I managed to crash my
} system 3 times in 2 hours so I wouldn't recommend it.

I am lucky enough to use both NIH-Image and Image Pro Plus for
windows, and both programs work perferct on my hardware. I wonder
which version of Image Pro Plus Dough used - the one (ver 1.1) I am
using seem to be OK.

I do not think it is fair to compare any program running two machines
as different as a high en MAC and mid range PC. I guess the
difference in your filtering would have been much smaller if you had
run Image Pro Plus on a highspeed Pentium based PC. But who really
cares about a few milli-seconds - I run most analysis as unatended
macros anyway. What counts for most users is that there image
analysis program works like the rest of their programs. So - if your
a MAC user download NIH-Image. If your a Windows user buy Image Pro
plus (or Mocha or...).
I am lucky to use both.

Bo


___________________________________________________________________
Bo Johansen E-Mail: BoJ-at-bot.ku.dk
Botanical Laboratory Vioce: +45 3532 2150
Gothersgade 140
DK-1123 Copenhagen K, Denmark
-------------------------------------------------------------------





From: Niels Christian Krieger Lassen 5791. :      AFM-NCKL-at-risoe.dk
Date: Wed, 12 Oct 1994 17:27:13 MET
Subject: subscribe afm-nckl@risoe.dk

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe afm-nckl-at-risoe.dk




From: David Hall :      hall-at-aecom.yu.edu
Date: Wed, 12 Oct 1994 13:56:28 -0400 (EDT)
Subject: more on image analysis: Bioquant

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We would be interested to hear from users of the Bioquant package,
particularly regarding the OS/2 version, and its suitability for doing
image analysis, cell tracing and measurements via camera lucida
from a light microscope.

Does the package perform well? Is it especially difficult to operate?
etc.

Thanks

Lucy Brown brown-at-aecom.yu.edu
Dept Neuroscience
Albert Einstein College of Med
Bronx, NY 10461








From: Fermin, Cesar :      Fermin.Cesar-at-tmc.tulane.edu
Date: 11 Oct 1994 17:31:39 -0600
Subject: Tissue bank protocol inquiry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




I would appreciate any information sent directly to me at
*Fermin-at-TMC.Tulane.edu* on guidelines that anyone may have on the operation of
a repository for frozen tissues to be used for research an to serve as an
intitutional resource. The information may also be faxed to me at the number
below. Thanks in advance for any information that I may use to establish
guidelines for this important research resource. Yes, imaging is done in some
of the tissues, so I am not stretching it too much by posting it here. Sorry!

************************************************************
*Cesar D. Fermin, Ph.D \|*|/ Fax (504) 587-7389 *
*Tulane Medical School /|*|\ Answ. Mach.(504) 584-261 *
*Pathology/SL79 \|*|/ Secretary (504) 584-2436 *
*New Orleans, La 70 112 /|*|\ Lab (504) 5841 *
*Fermin-at-TMC.Tulane.edu -} Director of Morphological Services*
************************************************************




From: Kerry Gascoigne :      mnklg-at-cc.flinders.edu.au
Date: Thu, 13 Oct 1994 16:16:48 +0000
Subject: Etec Umbilical

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410130700.AA26646-at-gamgee.cc.flinders.edu.au}
Sender: {mnklg-at-cc.flinders.edu.au}

I am having difficulty obtaining some 'coaxicon' miniature coaxial
connectors to make up a 'nimbin' extension cable to enable some
trouble shooting to occur on our Etec scanning microscope. The
Australian AMP agents just don't want to know about them. Can any one
recommend an AMP outlet in the USA that might have these items?
I need 8off partnumber 201143-1, 8off201144-1 and 16 0ff partnumber
328666-0.
Thanks,
Kerry Gascoigne
Research EM Unit,
Flinders University
South Australia




From: Doug Arrell :      ARRELL-at-jrc.nl
Date: Thu, 13 Oct 1994 10:23:42 GMT+0200
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941013102342.896-at-FS-IAM-1.JRC.NL}

} Doug Arrell wrote:
}
} } I know this sounds negative, but I wouldn't bother. My recomendation
} } would be to by a Macintosh and download NIH-Image from
} } ZIPPY.NIMH.NIH.GOV
} } as this will cost you roughly the same as a 'cheap' commercial
} } package on a PC and in my (admittedly limited) experience is
} } superior. This software is very flexible and gives you virtually
} } direct e-mail access to the author via a user group if you have any
} } problems. The software is also very fast (for a non-hardware analysis
} } solution). For example I tested Image Pro Plus on a 33Mhz 486 PC
} } doing a 'sharpen' filter and this woked at roughly 40000 pixels per
} } second, whereas NIH Image on a PowerPC 8100 (80Mhz) managed about
} } 710000 pix/sec.
} } Incidentally Image Pro Plus under windows has some bugs that I found
} } within 1 hr of starting to try it out, and I managed to crash my
} } system 3 times in 2 hours so I wouldn't recommend it.
}
} I am lucky enough to use both NIH-Image and Image Pro Plus for
} windows, and both programs work perferct on my hardware. I wonder
} which version of Image Pro Plus Dough used - the one (ver 1.1) I am
} using seem to be OK.
}
} I do not think it is fair to compare any program running two machines
} as different as a high en MAC and mid range PC. I guess the
} difference in your filtering would have been much smaller if you had
} run Image Pro Plus on a highspeed Pentium based PC. But who really
} cares about a few milli-seconds - I run most analysis as unatended
} macros anyway. What counts for most users is that there image
} analysis program works like the rest of their programs. So - if your
} a MAC user download NIH-Image. If your a Windows user buy Image Pro
} plus (or Mocha or...).
} I am lucky to use both.
}
} Bo
I used version 1.1 also. To get some strange results try doing some
pasteing of seperate images into one larger one. By doing this I
managed to put part of the image into the information line at the
bottom of the window, though how I did it I have no idea!

Doug Arrell


+------------------------------------+
| Dr Douglas Arrell |
| Mechanical Performance and Joining |
| Institute for Advanced Materials |
| 1755 ZG Petten |
| Netherlands |
| {ARRELL-at-JRC.NL} |
| Tel. (+31) 2246 5287 |
| Fax (+31) 2246 1917 |
+------------------------------------+




From: Doug Arrell :      ARRELL-at-jrc.nl
Date: Thu, 13 Oct 1994 10:33:40 GMT+0200
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941013103340.800-at-FS-IAM-1.JRC.NL}

} In reading your note to Joe Micheal about the NIH Image - I noted
} you used a PowerPC - is there a working native Power PC version of NIH out
} yet and / or do you have any comments about operating on the powerPC vs
} plain MAC. I just order a 7100 to run a Gatan camera system and frankly
} I'm worried about the PowerPC in non native applications.
}
This message was sent yesterday by Wayne Rasband (the author of NIH
Image).

} } } I have uploaded a new PowerPC beta that supports plug-ins. There
is now only one missing feature in the PPC native version: you can't
open or print documents from the Finder....

--wayne
} } } }

As far as I know, non-native applications work fine, just slower than
on a top-spec 'normal' mac, though any screen updating is faster.
However, what I am telling you is not from personal experience, I
have used NIH Image on various macs, but our new PowerPC has not
arrived yet. When it does I will give you some info.

Doug Arrell




+------------------------------------+
| Dr Douglas Arrell |
| Mechanical Performance and Joining |
| Institute for Advanced Materials |
| 1755 ZG Petten |
| Netherlands |
| {ARRELL-at-JRC.NL} |
| Tel. (+31) 2246 5287 |
| Fax (+31) 2246 1917 |
+------------------------------------+




From: Jean-Luc Rouviere :      rouvier-at-drfmc.ceng.cea.fr
Date: Thu, 13 Oct 1994 10:14:54 -0500
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X400-Received: by mta bottom.ctd.anl.gov in /PRMD=ANL/ADMD= /C=US/; Relayed;
Thu, 13 Oct 1994 08:15:21 -0500
X400-Received: by /PRMD=ESnet/ADMD= /C=US/; Relayed;
Thu, 13 Oct 1994 08:15:16 -0500
X400-Received: by /PRMD=inria/ADMD=atlas/C=fr/; Relayed;
Thu, 13 Oct 1994 08:15:09 -0500
X400-Received: by /PRMD=cea/ADMD=atlas/C=FR/; Relayed;
Thu, 13 Oct 1994 10:14:54 -0500

Several people are worried about NIH image on a Mac PowerPC. So do we, but we need a version of NIH image with FFT !
We would be interested to know if there is a version of NIH image with FFT that can work on a Mac Intosh Power PC (may be a non-FPU version ?). We tried to run an old version of image FFT on a Power Mac and it refused to work asking for a coprocessor...

PS. By the way, why the FFT menu has not been included in the standard NIH image program ?

Thanks very much for any information.





From: Doug Arrell :      ARRELL-at-jrc.nl
Date: Thu, 13 Oct 1994 15:55:31 GMT+0200
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941013155531.512-at-FS-IAM-1.JRC.NL}

} Problems with merging two smaller images into one larger sounds
} like a memory problem with the PC system. Just not enough there.
}
I thought so too, but I was working with two 200k images in 8Mb of
RAM + a swap file. My system reported 21Mb of free memory, so I don't
think that was the problem!

Doug Arrell


+------------------------------------+
| Dr Douglas Arrell |
| Mechanical Performance and Joining |
| Institute for Advanced Materials |
| 1755 ZG Petten |
| Netherlands |
| {ARRELL-at-JRC.NL} |
| Tel. (+31) 2246 5287 |
| Fax (+31) 2246 1917 |
+------------------------------------+




From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Thu, 13 Oct 1994 08:21:56 -0700 (PDT)
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm running NIH-Image on a PPC 8100. The 680x0 code version runs just
fine on the PPC. The question of speed is relative. Compared to the
MacIIcx I was using, the PPC is a screamer. If you are comparing to the
most recent Quadra and Centris machines, you won't be quite so
impressed. I've been testing the beta releases of Image in PPC native
code and the capability is absolutely amazing. haven't yet tried the
latest beta (.30) yet. Tlhe .28 release had problems so don't try it.

As far as low cost image analysis packages for the pc, doesn't Data
Translation offer an inexpensive basic package? I thought you could get a
price break if bundled with one of their frame grabber cards.

Regards,

Glen MacDonald
Hearing Development Laboratories RL-30
University of Washington
Seattle, WA 98195
(206)543-8360
glenmac-at-u.washington.edu






From: vaytek-at-INS.INFONET.NET
Date: Thu, 13 Oct 1994 11:08:32 CST
Subject: Re: Image Analysis Software for PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thank you Bo.

It is refreshing to see someone present an objective viewpoint. With a few
exceptions, most software and hardware products have strengths and weaknesses.
Performance is determined by clock speeds, CPU type (CICS - RISC), the amount
of RAM and CACHE memory, the BUS type, 16 bit / 32 bit operating system, etc...
Trying to compare two very different hardware/software configurations is a very
painstaking task that computer magazines spend a lot of ressources trying to do
with some kind of objectivity.

We all get used to what we have and use every day. Thanks Bo for not making a
recomendation based on your personal preferences.

Patrick Guerin
Vaytek, Inc.
Fairfield IA 52556
515-472-2227





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Fri, 14 Oct 1994 12:50:50 -0500 (CDT)
Subject: SuperSem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410131856.AA03243-at-esds01.es.dupont.com}


The program Super SEM, a supercard program for teaching principles
of SEM is being developed by the Microscopy Group at the
University of Western Australia in Perth. You should contact
Brendon Griffin at the following address for more details. Use his
UWA address (bjg-at-uniwa.uwa.edu.au) as his mail is forwarded and following
him around the world, while he is on sabbatical leave..

Brendon J. Griffin
Visiting Research Fellow Until 10th Nov., 1994
Crystallography & Mineral Physics Unit NB email address is always
Department of Geological Sciences current
University College of London Ph: 071-380-7777 Ext 2406
Gower Street Fax.: 071-388-7614
London WC1E 6BT Email: bjg-at-uniwa.uwa.edu.au


----

Nestor




From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Fri, 14 Oct 1994 14:16:29 -0400 (EDT)
Subject: RE: Inexpensive Light Microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Fri, 14 Oct 1994 13:00:07 -0400, gnord-at-mactem.er.usgs.gov writes:

} Hello Microscopists, especially the photon variety.
}
} I am looking for a light microscope for about $150 for a seven year old
} girl very interested in science. Zany-Brainy, a children's learning
} center store here in Virginia, has a toy for about $70. I think that I
} can do better. Any ideas for new or used light microscopes in my price
} range?
} Gordon L. Nord Jr.
} 959 National Center
} U. S. Geological Survey
} Reston, VA 22092
}
} Office: 703-648-6745
} FAX: 703-648-6789
}
} gnord-at-mactem.er.usgs.gov
}
===============================
Edmund Scientific, 101 E. Gloucester Pike, Barrington, NJ 08007-1380, Tel.
order 609 573 6250 (Schools & Ed. pricing 609 573 6270) sells "student"
microscopes in your price range.

*****************************************************
M.V. Parthasarathy
Section of Plant Biology, 228 Plant Science Building
Cornell University, Ithaca, NY 14853. USA
Tel: 607-255-1734; Fax: 607-255-5407
E-Mail: mvp2-at-cornell.edu
******************************************************
e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-
e- e-
e- This message was sent using 100% recycled electrons e-
e- e-
e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-




From: Michael Boucher :      BOUCHER-at-tcrca.usbm.gov
Date: Fri, 14 Oct 1994 13:26:19 CST
Subject: SEM-need advice on surplusing or junking

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have an old JEOL model JSM-2 SEM with a Nuclear Diodes/ EDAX
model 508 EDS system. It was built in 1968. Although in good
operating condition, we have no use for it and I have suggested that
probably no one would want it. I need to somehow verify that no gov't
agency, or University would want to acquire it by coming here,
packing it up and shipping it. Further that no citizen would bid on
it and also come to Mpls to get it. Is there a market or demand for
old SEMs or will we have to pay someone to scrap it for metal? Any
comments can be sent to me at the e-mail address below.
Thanks for your input.
Mike Boucher Boucher-at-tcrca.usbm.gov
*****************************************************************
Michael L. Boucher Sr. Boucher-at-TCRCA.USBM.GOV
Geology-Mineralogy/Chemistry Labs Ph 612-725-4614
Twin Cities Research Center Fax 612-725-4526
U.S. Bureau of Mines Center 725-4500
Department of Interior
5629 Minnehaha Avenue South
Minneapolis, MN 55417-3099
U.S.A.
*****************************************************************





From: ARGIL-at-delphi.com
Date: Fri, 14 Oct 1994 22:35:09 -0400 (EDT)
Subject: inexpensive microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Gordon Nord wrote:

} I am looking for a light microscope for about $150 for a seven year old
} girl very interested in science. Zany-Brainy, a children's learning
} center store here in Virginia, has a toy for about $70.
} I think that I can do better.

Hello Gordon -

My company imports all types of microscopes. We have a model for your price
range that has 40X - 640X, 3 sets of eyepieces, mirror illuminator with
aperture disc, wooden box. It is very heavy, and the optics are
excellent. An electric substage illuminator is available for a little extra.
We have even used this one with video, and it works very well.

I would be happy to discuss your needs. Send me a note for more info.

Arthur Gillman
Princeton, NJ




From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 15 Oct 1994 00:06:19 U
Subject: FACULTY POSITION

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1429942656.73704-at-mse.engin.umich.edu}

John Mansfield
North Campus Electron Microbeam Analysis Lab
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143 Phone: (313)936-3352
jfmjfm-at-engin.umich.edu or
John.F.Mansfield-at-umich.edu Time:
12:07 AM

Date:10/15/94
NC EMAL

FACULTY POSITION

THE UNIVERSITY OF MICHIGAN

The Department of Materials Science and Engineering at The University of
Michigan is continuing to grow and there are openings for tenure track
faculty positions at the level of assistant professor. Candidates at the
senior level with records of outstanding accomplishment will also be
considered. Applicants must have a Ph.D. degree, be qualified to teach
undergraduate and graduate courses in materials science and engineering, and
should plan to develop independent and cooperative research programs. A
demonstrated research record or potential is required. Preferred areas of
expertise include: electronic, magnetic and/or photonic materials; physical
metallurgy; materials theory and modeling; ceramics. Women and minority
candidates are encouraged to apply.
Send curriculum vitae and a list of references to:

Chairman, Search Committee
Department of Materials Science and Engineering
The University of Michigan
2300 Hayward Street
Ann Arbor, MI 48109-2136

An Equal Opportunity Affirmative Action Employer







From: Philips Electron Optics :      MARCOM-at-eo.ine.philips.nl
Date: 17 Oct 94 17:50:26 GMT
Subject: TEM workshop on Convergent Beam Electron Diffraction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-99.941017175027.448-at-eo.ine.philips.nl}
To: microscopy-at-aaem.amc.anl.gov

PHILIPS WORKSHOP ON CONVERGENT BEAM ELECTRON DIFFRACTION

March 27-31, 1995


PURPOSE

The purpose of this workshop is to instruct microscopists in the
potential of CBED and large angle CBED for the analysis of crystal
structures, and to provide hands-on training in the practical
techniques as well as example classes dealing with various
applications. Although some background in diffraction techniques
might help, it is not absolutely necessary and a quick refresher
course on diffraction theory will start the week.


TOPICS

The following questions will be answered:
1) What information is contained in a CBED pattern?
2) How are crystal symmetries determined?
3) How are CBED and LACBED done in practice?
4) What are the practical and instrumental limitations?
5) How can crystal defects be analysed?


INSTRUCTORS AND EQUIPMENT

The teachers (from renowned convergent beam groups such as Bristol,
UK; Lille, France and Perth, Australia) have a wide experience in
organising workshops in this field and will be assisted by Philips
application specialists. Invited lectures on more or less advanced
CBED applications will be presented by specialists such as Prof. J.P.
Morniroli, Dr. D. Cherns and Dr. A. Johnson. A wide range of Philips
CM transmission electron microscopes (LaB6 & FEG) will be available to
perform all types of convergent beam experiments including energy
filtering of the diffraction patterns. Simulations and interpretations
of the experiment will be done on stand-alone systems.


MORE INFORMATION

The workshop will be held at the Philips Electron Optics Applications
Laboratory in Eindhoven, The Netherlands. Persons who wish to
register for the workshop or obtain more information should
send a reply.
Please enclose your complete postal address and indicate if you want
to register and/or receive more information.

With Kind Regards

Eric van Cappellen


FOR MORE INFORMATION CONTACT:

Philips Electron Optics
Building AAE, P.O. Box 218
5600 MD Eindhoven
The Netherlands
Tel. +31 40 766234
Fax. +31 40 766102
E-mail: marcom-at-eo.ine.philips.nl




From: Anna Carlsson :      OO2ANNA-at-robin.mbfys.lth.se
Date: Tue, 18 Oct 1994 11:42:00 +0100
Subject: Conference on EM in Solid State Science

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


EUCHEM CONFERENCE ON ELECTRON MICROSCOPY IN SOLID STATE SCIENCE
MAY 20-24, 1995 in LUND, SWEDEN

A EUCHEM conference on Electron Microscopy in Solid State Science will be held
at Jaravallen, at the seaside 20 km northwest of the university town Lund, on
May 20-24, 1995. In the spirit of the EUCHEM conferences, which are European
equivalents to the Gordon conferences in the US, there will be a limited
number of oral presentations, but plenty of time for discussions. No abstracts
or proceedings will be published. The number of participants will be limited
to 75 and all participants are requested to present a poster.

Among the topics that will be adressed are:
Electron microscopy of new materials.
Structure and chemistry of aperiodic features in solids.
New microscopy techniques and their application.

For further information, please contact:
The Swedish National Committee for Chemistry
Wallingatan 24, 3 tr
S-111 24 Stockholm
Sweden
Phone: Int +46-(0)8-4115280, Fax: Int +46-(0)8-106678
E-mail: Anna.Carlsson-at-oorg2.lth.se




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Tue, 18 Oct 1994 9:31:43 -0500 (CDT)
Subject: Quantitative N analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Alfred

If you want to do quantitative analysis of N in high Z materials you
better get a standard as close as possible to your "unknowns" the
matrix and absorption corrections will be significant. Using Si3N4
as your standard would not be a good idea. Sorry but I can't give
you any leads as to standards. Your best bet might be to consult
a phase diagram and make a standard.

Nestor




From: jmcgee-at-lunatic.er.usgs.gov (Jim McGee)
Date: Tue, 18 Oct 1994 10:56:39 -0400
Subject: nitrogen

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Alfred:

Geller Microanalytical (508 535-5595) has several nitride standards, including
a TaN. I don't know whether or not they are good, but I would hope so since they
are advertised and sold as standards. Bastin has numerous articles in the MAS
literature regarding N analysis (Microbeam Analysis 1988, p290, for example).
I am dabbling with N analyses in volcanic tuffs (appears to be in the form of
NH4 in feldspars - much lighter stuff than you are studying). Count rates are
poor, to say the least. Good luck.

Jim McGee



James J. McGee email: jmcgee-at-lunatic.er.usgs.gov
U.S. Geological Survey Phone: (703) 648-6782
959 National Center Fax: (703) 648-6789
Reston, VA 22092





From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Tue, 18 Oct 1994 10:55:10 -0500 (CDT)
Subject: paraffin knife holder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Greetings!

I have a very old (wrought iron) A/O Spencer paraffin microtome Model 820
with a knife holder for large reusable steel blades. Our lab would like
to purchase (or trade) a disposable knife holder that would fit our
machine.

Suggestions for where to look would also be appriciated!

Thanks,
Kathy Walters
CEMRF
University of Iowa






From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Tue, 18 Oct 1994 11:04:36 -0500 (CDT)
Subject: nonvender service contracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi again,

Does anyone know of a service company in my area that does quality
repair and preventative maintanance on EMs?

More specifically we have 4 Hitachi microscopes, 2 SEMs (S2700 &
S4000 with FE) and 2 TEMs (H-600 & H-7000).

Thanks,
Kathy Walters
CEMRF
University of Iowa
Iowa City






From: david.rayns-at-stonebow.otago.ac.nz (David Rayns)
Date: Wed, 19 Oct 1994 08:44:04 +1200
Subject: cytoskeleton

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199410182031.AA10353-at-arwen.otago.ac.nz}
X-Sender: st004084-at-brandywine.otago.ac.nz
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I wish to subscribe to the microscopy listserver.
I am a structural cell biologist with a major interest in cytoskeleton. My
main tools are TEM, various freeze-preparation methods and tissue culture.
I am very interested in exchanging ideas on the functional ultrastructure
of cytoskeleton. For example, can anyone suggest a suitable 3hr practical
class execise for third year BSc students based on microtubules including
at least some live material.
Thanks!






From: Daniel Possin :      oemlab-at-u.washington.edu
Date: Tue, 18 Oct 1994 13:40:01 -0700 (PDT)
Subject: LR White LM section mounting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: oemlab-at-stein3.u.washington.edu

Hello everyone -

Does anyone out there care to share their sure-fire way to securely mount
LR White 1-2 micron sections to glass slides for immunogold labeling? We
have variable success with keeping our sections from bubbling, wrinkling
and working their way loose during immuno- incubations. Any secrets you
would care to post would be much appreciated.

Thanks in advance.

Dan Possin
Ophthalmology EM Lab, Univ. of Washington, Seattle, USA.





From: Daniel Possin :      oemlab-at-u.washington.edu
Date: Tue, 18 Oct 1994 14:37:03 -0700 (PDT)
Subject: LR White LM section mounting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: oemlab-at-stein2.u.washington.edu

Hello everyone -

Does anyone out there care to share their sure-fire way to securely mount
LR White 1-2 micron sections to glass slides for immunogold labeling? We
have variable success with keeping our sections from bubbling, wrinkling
and working their way loose during immuno- incubations. Any secrets you
would care to post would be much appreciated.

Thanks in advance.

Dan Possin
Ophthalmology EM Lab, Univ. of Washington, Seattle, USA.




From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Tue, 18 Oct 1994 16:55:15 -0500
Subject: slide coating for LM sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Hello everyone -
}
} Does anyone out there care to share their sure-fire way to securely mount
} LR White 1-2 micron sections to glass slides for immunogold labeling? We
} have variable success with keeping our sections from bubbling, wrinkling
} and working their way loose during immuno- incubations. Any secrets you
} would care to post would be much appreciated.
}
} Thanks in advance.
}
} Dan Possin
} Ophthalmology EM Lab, Univ. of Washington, Seattle, USA.

After trying a zillion different coatings (e.g., poly-l-lysine, chrome-gel,
albumin), we now use aminopropy-triethoxysilane (Sigma). Dip clean slides
in a 2% apts in acetone solution for 30 sec. rinse in acetone, then dH20.
dry and store. stable for at least months. this protocol is based on one
of Lynne Angerer that she present at a ASCB in situ hybridization workshop.
she sites Gottlieb and Glaser (BBRC 63:815-821, 1976) who used apts coated
slides treated with glutaraldehyde to adhere single cells. she states that
glut post treatment is good for cells but unnecessary for sections. our
experience is in agreement with this. Additional advantage: Water beads
up on these slides so that when one adds 10-20 ul of antibody, it doesn't
spread flat on the slide but stays restricted to a bead.

good luck

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Tue, 18 Oct 1994 22:21:07 -0700 (PDT)
Subject: Re: LR White LM section mounting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer03.u.washington.edu

I've had to resort to slides washed in chromic acid and then subbed with
chrome alum-gelatin for some combinations of plastics and mountants. We
have a bunch of other slide treatments for specific, and general,
purposes. Drop in and try a few. Dale can get some for you if I'n not
around.

Glen MacDonald
Hearing Development Laboratories RL-30
University of Washington
Seattle, WA 98195
(206)543-8360
glenmac-at-u.washington.edu



On Tue, 18 Oct 1994, Daniel Possin wrote:

} Hello everyone -
}
} Does anyone out there care to share their sure-fire way to securely mount
} LR White 1-2 micron sections to glass slides for immunogold labeling? We
} have variable success with keeping our sections from bubbling, wrinkling
} and working their way loose during immuno- incubations. Any secrets you
} would care to post would be much appreciated.
}
} Thanks in advance.
}
} Dan Possin
} Ophthalmology EM Lab, Univ. of Washington, Seattle, USA.
}




From: Dr R.J. Keyse :      keyse-at-liverpool.ac.uk
Date: Wed, 19 Oct 1994 11:52:24 +0100 (BST)
Subject: AEM-Nitrogen & M-lines

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410190452.AA35318-at-pukrs7.puk.ac.za}

The nitrogen-K emission line is at 392eV, there are many M zeta lines
for elements in the mid-high atomic number range there. Tin for example
is at 397eV. These (Mz) lines are however fairly weak!

By interpolating between elements I would say molybdenum ought to
have a M-gamma (strongish line) at around the N-k that could cause a problem.
This line is not listed for some reason! All I find for Mo M-series is M3-N1
at 331eV (strength listed as 100).

In the tables I use for X-ray emission line identification with EDX
there's no listing for indium M-lines. (Johnson and White, ASTM Data
series DS46, 1970). My EDX computer system likewise cannot show the
M-lines for indium. Below In is Cd, Ag, Pd down to Nb, all of which
have M lines with M-gamma the strongest. With the exception of Mo noted.
Indium has auger and XPS peaks for M transitions, there's an EELS edge too.

Is there another (better) reference to use for the M-series?

Robert Keyse (keyse-at-liverpool.ac.uk)




From: MicroToday-at-aol.com
Date: Wed, 19 Oct 1994 15:02:01 -0400
Subject: Microscopy Today

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Contributers to our newsletter "Microscopy Today" please note our new eMail
address - MicroToday-at-aol.com.
Others, including international, who wish no cost subscriptions, have but to
supply their complete mailing addresses by eMail - or by fax (608/836-1969)




From: Charles E. Lyman :      cel1-at-Lehigh.EDU
Date: Wed, 19 Oct 1994 16:46:45 -0400
Subject: M-lines

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bob Keyse,

Some of the lines you listed I never see. One of the best sources for x-ray
lines is the Chemical Rubber Co. (CRC) Handbook of Chemistry and Physics. This
contains the famous Bearden tables of x-ray wavelengths and energies. You are
right about In M-lines. Even in this table (mine is dated 1967) the low energy
indium lines are incomplete; perhaps a more up-to-date version would have them.

Basically there are five M-lines that may be observed in the EDS spectrum. We
usually only see the two strongest M-alpha and M-beta (typically unresolved).
The other three (zeta, gamma, and M2N4) are only about 5% or less of the
M-alpha.

I obtained the latter information from Table 3.11 of Goldstein et al., Scanning
Electron Microscopy and X-ray Microanalysis, Plenum Press, New York, 1992, p.
127. Chapter 6 on qualitative analysis shows typical spectra for all types of
lines from various elements: for example, M-lines from Bi, Ta, and Dy.

Good luck, Charlie

Prof. Charles E. Lyman
Department of Materials Science and Engineering
Lehigh University, 5 East Packer Avenue, Bethlehem, PA 18015-3195
E-mail: cel1-at-Lehigh.edu Tel: 610-758-4249 FAX: 610-758-4244





From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Thu, 20 Oct 1994 12:05:31 +1300
Subject: Re equipment list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0qxiw7-000fIbC-at-pegasus.cc.ucf.edu}

}
} To:Robert McDonald {robert-at-geology.gla.ac.uk}
} From:richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
} Subject:Re equipment list
}
}
} } On another note - do the other folks reading this group think that
} } would be useful to compile a list of what equipment we all have
} } so that like users could pool info on specific problems etc?
} }
} } Just a thought and my opinions only.
} }
} } Robert McDonald robert-at-starav.geology.gla.ac.uk
}
} I think Robert's idea re equipment list is a good idea but would take an awful
} lot of work, compiling, updating etc as I am sure that a list of just the EMs
} used by the people who use this listserver would be frighteningly long.
} Perhaps it would be better if those who need help with specific pieces of
} equipment post their problems on the listserver and just keep the addresses of
} respondants for future reference, provided they are happy to be contacted
} again for assistance.
} I don't deny it would be interesting to see what's out there though.
}

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Thu, 20 Oct 1994 12:05:31 +1300
Subject: Re equipment list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {1994Oct19.164942.1515942618-at-ms.sjdccd.cc.ca.us}
To: MICROSCOPY-at-aaem.amc.anl.gov (MSA list)

MSA, Microscopy Soc of Am does have a list of facilities with their
equipment collected by the Tech Forum. They have it on disk.
contact
Sandy Silvers
EM Complex
USDA,ARS,RRC
PO Box 5677
Athens, GA 30613-6199
706/546-3471
No e-mail address listed.

I collected all the training facilities in the country and in several other
countries with all equipment listed, but more emphasis on the course info.
This however is outdated since I did this crazy thing in 1982. It was a
marvelous amount of information but a fantastic amount of work!!!!
Good Luck
Judy M

Judy Murphy, PhD
Dept. of Microscopy
San Joaquin Delta College
5151 Pacific Ave
Stockton, CA 95207

Phone: 209/474-5284
FAX: 209/474-5649
e-mail:
murphy.ms.sjdccd.cc.ca.us

_______________________________________________________________________________


}
} To:Robert McDonald {robert-at-geology.gla.ac.uk}
} From:richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
} Subject:Re equipment list
}
}
} } On another note - do the other folks reading this group think
that
} } would be useful to compile a list of what equipment we all have
} } so that like users could pool info on specific problems etc?
} }
} } Just a thought and my opinions only.
} }
} } Robert McDonald robert-at-starav.geology.gla.ac.uk
}
} I think Robert's idea re equipment list is a good idea but would take an
awful
} lot of work, compiling, updating etc as I am sure that a list of just the
EMs
} used by the people who use this listserver would be frighteningly long.
} Perhaps it would be better if those who need help with specific pieces of
} equipment post their problems on the listserver and just keep the
addresses of
} respondants for future reference, provided they are happy to be contacted
} again for assistance.
} I don't deny it would be interesting to see what's out there though.
}

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301










From: Richard E. Edelmann :      REDELMAN-at-musom01.mu.wvnet.edu
Date: Thu, 20 Oct 1994 07:57:03 +1100
Subject: Re: M-lines

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Brief addition to Charles E. Lyman's suggestion of looking in the
rubber book (HB Chem & Physics): I just looked in my 1984-1985 copy
and the low energy Indium M lines are not present either.

Richard E. Edelmann
Electron Microscopy Facility Supervisor
Marshall University - School of Medicine
Huntington, West Virginia




From: gkennedy-at-ucsd.edu
Date: Thu, 20 Oct 1994 08:40:56 -0800
Subject: LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199410201537.IAA09536-at-ucsd.edu}
X-Sender: gkennedy-at-popmail.ucsd.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
Content-Transfer-Encoding: quoted-printable

Last week, someone asked about LR White sections/semi-thins on slides for=
immuno. I have not had this problem with my material, BUT I am meticulous=
when I clean my slides before applying any type of slide adhesive. I use=
ordinary gelatin/chrome alum subbing solution (0.5% gelatin, ordinary 300=
bloom, and 0.05% chromium potassium sulfate). I thoroughly clean the=
slides with a good grease-cutting detergent, then rinse with hot tap H2O,=
then de-i H20, then dip. I suggest this as a substitute for acid-cleaning,=
as it's a little safer. I must caution that the rinsing step is=
critical-any soap residue is will render the slides useless (we had a=
technician who failed to follow directions and caused a few major=
catastrophies...). In closing, I'd like to thank everyone who responded to=
my cry for help with LR White thin sections. I think my resin was a little=
too elderly. Grace






From: lelia-at-srmp04.saclay.cea.fr
Date: Thu, 20 Oct 1994 11:50:44 -0500
Subject: Allied Technology address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X400-Received: by mta bottom.ctd.anl.gov in /PRMD=ANL/ADMD= /C=US/; Relayed;
Thu, 20 Oct 1994 11:52:58 -0500
X400-Received: by /PRMD=ESnet/ADMD= /C=US/; Relayed;
Thu, 20 Oct 1994 11:52:56 -0500
X400-Received: by /PRMD=inria/ADMD=atlas/C=fr/; Relayed;
Thu, 20 Oct 1994 11:52:51 -0500
X400-Received: by /PRMD=cea/ADMD=atlas/C=FR/; Relayed;
Thu, 20 Oct 1994 11:50:44 -0500

I'm trying to get in toch with Allied Technology but have not yet
succeeded in getting their coordinates.
If anyone can send their address, telephone and/or fax numbers,e-mail
it will be greatly appreciated

Lelia Schmirgeld-Mignot

---------------------------------------------------------------------
SRMP/DECM/DTA Tel : 33-1-69.08.20.68
C.E. Saclay FAX : 33-1-69.08.68.67
91191 GIF sur YVETTE Cedex e-mail : lelia-at-srmp04.saclay.cea.fr
FRANCE
---------------------------------------------------------------------





From: images1-at-biosci.mbp.missouri.edu (Michael Stanley)
Date: Thu, 20 Oct 1994 14:19:22 -0500
Subject: diaphot prism

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Folks,
We have a Nikon Diaphot connected to our Bio-Rad Confocal. The
slider on the prism that shunts the laser up or the transmitted light to
the oculars is getting very hard to pull. Nikon originally said this is
natural and to replace the silicon gel on the prism glides inside the
housing, preferably by their own gel. They now say that we should use a
high quality oil instead of the gel.
Since this is a big deal, hard to do thanks to the confocal
assembly, I thought I would ask if anyone else out there has run into this
problem.
TIA

C. Michael Stanley
Molecular Cytology Core Facility
Molecular Biology
2 Tucker Hall
University of Missouri-Columbia
Columbia, MO. 65211
(314) 882-4895
fax= 314-882-0123






From: Laurie Frederick :      frederick_laurie-at-vanlab.paprican.ca
Date: Thu, 20 Oct 1994 13:20:43 PDT
Subject: Re: Allied Technology address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941020132043.416-at-vanlab.paprican.ca}
To: lelia-at-srmp04.saclay.cea.fr

} I'm trying to get in toch with Allied Technology but have not yet
} succeeded in getting their coordinates.
} If anyone can send their address, telephone and/or fax
numbers,e-mail
} it will be greatly appreciated
}
} Lelia Schmirgeld-Mignot

Hi,
I have an address for an E.M. supplier called
Allied Hi-Tech Products,
2376 E. Pacifica Place
Rancho Dominguez, CA 90220

There numbers are:
Phone (800) 950-9347
Fax (310) 762-6808

Hope this helps,
Laurie

______________________________________________________________________
Laurie Frederick, A.SC.T. PAPRICAN
Corrosion Control Group 3800 Wesbrook Mall
The Pulp and Paper Research Vancouver, B.C.
Institute of Canada Canada V6S 2L9

Email: frederick_laurie-at-vanlab.paprican.ca
Tel: 604-222-3200 Fax: 604-222-3207




From: XiaoGuang Ning :      ningx-at-mcmail.cis.mcmaster.ca
Date: Thu, 20 Oct 1994 17:11:51 +0059 (EDT)
Subject: information request

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Manager:

Would you please tell me how to subscribe MICROSCOPY?

I subscribed by command "subscribe microscopy
ningx-at-mcmail.cis.mcmaster.ca" the MICROSCOPY IN
"LISTSERVER-at-ANLEMC.MSD.ANL.GOV" yesterday, but no respond.

With Best Regards


Dr.Xiaoguang NING






From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Thu, 20 Oct 1994 17:30:08 -0400 (EDT)
Subject: Re: diaphot prism

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} the oculars is getting very hard to pull. Nikon originally said this is
} natural and to replace the silicon gel on the prism glides inside the
} housing, preferably by their own gel. They now say that we should use a
} high quality oil instead of the gel.

We have had the same problem on all our Diaphots.
Nikon claims that their new oil is especially heat resistant and far better
than past lubricants.
They have cleaned and relubricated the Diaphot we are using with the
confocal and it is now much easier to use. We plan to have one of the
other Diaphots greased the same way.
We have an additional problem with the prism which is looseness on its
track, but this is probably unrelated to the new lubricant unless the
machining got worn over the years, perhaps when it was tighter.
-Michael Cammer






From: POSHEL-at-wpo.it.luc.edu
Date: Fri, 21 Oct 1994 08:21:20 -0500
Subject: uProbe-Colloidal silver. -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Leon Smuts,
I've found that silver paints and pastes *seem* to dry quickly, but
actually degass for several hours, or a day or two, even using a vacuum
dessicator.
I've also found some brands of carbon paint that are contaminated with
phosphorus. Don't recall which, off the top of my head, and the
formulation have been changed, but it's something that should be
checked, especially if you're doing EDX or WDX.
Phil Oshel
poshel-at-luc.edu





From: POSHEL-at-wpo.it.luc.edu
Date: Fri, 21 Oct 1994 08:22:39 -0500
Subject: Colloidal Graphite -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0qyK7m-000fIbC-at-pegasus.cc.ucf.edu}

Re: Robert McDonald's question-- I agree that an equipment list would be
useful.
Phil Oshel
poshel-at-luc.edu





From: Ciara Mullan :      mullanc-at-mcmail.cis.mcmaster.ca
Date: Fri, 21 Oct 1994 12:15:41 +0059 (EDT)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0qyLyE-000fIbC-at-pegasus.cc.ucf.edu}



On Fri, 21 Oct 1994, Stephen Anderson wrote:

} Subject: Etchant for InGaAs. } } G'day ... } } A colleague of mine needs
to find a suitable etchant for InGaAs (indium } content x { 0.3), which
will reveal decorated defects inside the material } when the etched
surface is examined with Nomarski IC microscopy. } } Any comments or
references? } } Stephen Anderson. }
.............................................................. } : Stephen
Anderson : } : Electron Microscope Unit : } : The University of Sydney
Email stephen-at-emu.su.oz.au : } : NSW 2006 Telephone (+61)-2-351 2351 : } :
Australia Facsimile (+61)-2-552 1967 : }
:............................................................: }

A weak solution of Cl2 bubbled through a beaker of methanol will do it-
be careful if the solution becomes saturated,ie a green-yellow colour it
may ignite-have some extra methanol beside you if you try this. More
esoterically a colleague of mine suggested a solution of KOH:
K3Fe(CN)6:H2O in the ratio 6g:4g:50ml. He uses it for SEM imaging of
epilayers and reckons that the etchant rate is approximately 2
micrometers/minute at room temperature
Good luck
Ciara





From: Shu-Chun Su :      su-at-marlin.ssnet.com
Date: Fri, 21 Oct 1994 12:21:27 -0400 (EDT)
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe su-at-ssnet.com

******************************************************
* Shu-Chun Su *
* Hercules Incorporated Phone: 302-995-3498 *
* Research Center 8136-267 Fax : 302-995-4135 *
* 500 Hercules Road e-mail: su-at-ssnet.com *
* Wilmington, DE 19808-1599 *
******************************************************





From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Fri, 21 Oct 1994 18:21:52 -0400 (EDT)
Subject: RE: cytoskeleton

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Tue, 18 Oct 1994 16:44:04 -0400,
david.rayns-at-stonebow.otago.ac.nz (David Rayns) writes:

} I wish to subscribe to the microscopy listserver.
} I am a structural cell biologist with a major interest in cytoskeleton. My
} main tools are TEM, various freeze-preparation methods and tissue culture.
} I am very interested in exchanging ideas on the functional ultrastructure
} of cytoskeleton. For example, can anyone suggest a suitable 3hr practical
} class execise for third year BSc students based on microtubules including
} at least some live material.
} Thanks!
}
==============
David,

I can suggest a few lab excercises but they would require a fluorescence
microscope:

1. If you have monolayer cultured cells such as fibroblasts, you can
demonstrate the interrelatioship between microtubules (MTs) and ER
organization. The live cells can be stained with DioC6 to demonstrate the
normal ER organization and treatment of the cells with nocodazole will
result in the disorganization of ER (see Terasaki et. al. 1986, J. Cell
Bio).

2. If you have dividing cells, the importance of MTs
for chromosomal movement can be demonstrated using MT disruptants. You also
can include immunolocalization of the MTs and DAPI staining
of the chromosomes as part of the lab using fixed cells.

3. You can use onion scale epidermal cells to demonstrate the
importance of F-actin for protoplasmic streaming. Treating the cells with
cytoclasin D will stop the streaming. The presence or the lack of F-actin
can be determined by rhodamine-phalloidin (Rh-Ph) staining. (For details see
on Rh-Ph staining see Parthasarathy, 1987, Plant. Mol. Biol Rep.; Sonobe, S. and Shibaoka, H.,
1989, Protoplasma).

Hope this helps.




From: Jim Stanley :      jstanly-at-mse.ogi.edu
Date: Fri, 21 Oct 1994 15:25:15 -0700 (PDT)
Subject: Re: Re equipment list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I believe MSA has a relatively updated list of facilities through one of
their divisions.

On Thu, 20 Oct 1994, Richard Easingwood wrote:

} }
} } To:Robert McDonald {robert-at-geology.gla.ac.uk}
} } From:richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
} } Subject:Re equipment list
} }
} }
} } } On another note - do the other folks reading this group think that
} } } would be useful to compile a list of what equipment we all have
} } } so that like users could pool info on specific problems etc?
} } }
} } } Just a thought and my opinions only.
} } }
} } } Robert McDonald robert-at-starav.geology.gla.ac.uk
} }
} } I think Robert's idea re equipment list is a good idea but would take an awful
} } lot of work, compiling, updating etc as I am sure that a list of just the EMs
} } used by the people who use this listserver would be frighteningly long.
} } Perhaps it would be better if those who need help with specific pieces of
} } equipment post their problems on the listserver and just keep the addresses of
} } respondants for future reference, provided they are happy to be contacted
} } again for assistance.
} } I don't deny it would be interesting to see what's out there though.
} }
}
} Richard Easingwood
} Department of Anatomy and Structural Biology,
} P.O. Box 913
} University of Otago,
} Dunedin, New Zealand.
} Fax:64-3-479 7254
} Telephone:64-3-479 7301
}
}







From: knoff-at-lipovx.lbl.gov (Laura Knoff)
Date: Fri, 21 Oct 1994 15:25:15 -0700 (PDT)
Subject: MSA Undergraduate Research Scholarships

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: Microscopy-at-aaem.amc.anl.gov

Microscopy Society of America is accepting applications from undergraduate
students interested in conducting a research project involving the use of
ANY microscopy technique. Applicants must be sponored by a member of MSA.
The maximum award is $2500(US). The application deadline has been extended
to December 30, 1994. Applications can be obtained by contacting the MSA
Business Office at (800) 538-3672, FAX: (508) 548-9053.
For more information contact either:
Dr. Ralph Albrecht,rma-at-ahabs,wisc,edu (608) 263-3952/4162, FAX (602)
262-7420 ; or Dr. Richard Ornberg, rlornb-at-ccmail.monsanto.com (314)
694-1184, FAX (314) 694-6727.






From: Dwight Beebe :      beebed-at-ERE.UMontreal.CA
Date: Sat, 22 Oct 1994 17:46:39 -0400 (EDT)
Subject: TEM: looking for a book

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
I'd like to know whether anyone knows where I could purchase a
copy of the book by G. Meeks: "Practical Electron Microscopy for Biologists".
It is out of print and I'd like to find a copy somewhere. If anyone can
help, please reply directly to me, not to the list.
Many thanks,
Dwight

Dwight Beebe
IRBV, Dept. de sciences biologiques beebed-at-ere.umontreal.ca
Universite de Montreal Voice:514-872-4563
4101, rue Sherbrooke est FAX:514-872-9406
Montreal, PQ H1X 2B2 Canada







From: RETEP-at-anat.uct.ac.za
Date: 24 Oct 94 08:30:13 SAST-2
Subject: RE: Instrument List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A penny's worth for consumption.

I know that the majority of members of the listserver are in the
states and hence most of the communication on this subject has been
stateside orientated. However as an overseas 'consumer' of the
listserver it might be interesting to know of other lists in other
countries.

Certainly their is a list available in South Africa which is run by
Dane Gernecke and the UCT EM unit for EMSSA (Electron Microscopy
Society of Southern Africa).

Are there any others?

Peter
_______________________________________________________________


-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
-at- -at-
-at- Peter D. G. Richards -at-
-at- Dept Anatomy and Cell Biology -at-
-at- UCT Medical School -at-
-at- Observatory -at-
-at- 7925 -at-
-at- RSA -at-
-at- Tel: 021-406 6285. -at-
-at- Internet: retep-at-anat.uct.ac.za -at-
-at- -at-
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-




From: JJMILL-at-bunyip.ph.rmit.oz.au
Date: Mon, 24 Oct 1994 18:28:17 EST-10
Subject: list of instruments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941024182817.320-at-bunyip.ph.rmit.edu.au}
To: microscopy-at-aaem.amc.anl.gov

Re : recent topic

Australia has had for many years a full list of instrumentation and
personnel in EM maintained voluntarily by people from the national
association (ASEM). Te list is updated every two years by the
"owners" of the instrumentation and labs in conjunction with the
biennial EM conference. The list is a very valuable reference for
advice and all sorts of other things.
It would be very useful to have an international list, but would take a
while to arrange. Maybe this could be started via this server, for all to
list their instruments/people/interests in a tight format ??
jjm


Professor John J. Millar
Head, Department of Applied Physics
Electron Microscope Unit
RMIT, GPO Box 2476V, MELBOURNE AUSTRALIA 3001
+613 660 2602 fax 613 660 3837
email jjmill-at-rmit.edu.au




From: CAROLYN J. EMERSON, DEPT. OF BIOLOGY, MEMORIAL UNIVERSITY
Date: Mon, 24 Oct 1994 08:46:22 -0230
Subject: lists of instruments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: cemerson-at-kean.ucs.mun.ca
{cemerson-at-kean.ucs.mun.ca}
Reply-To: cemerson-at-kean.ucs.mun.ca
To: MICROSCOPY-at-AAEM.AMC.ANL.GOV
Message-Id: {009866A4.2FE243A2.1441-at-leif.ucs.mun.ca}

The Microscopiccal Society of Canada also has a Directory of Electron
Microscopy Facilities in Canada, last printed in 1992. We hope to have
an update in 1995 including confocal and scanned probe instruments as well
Directories are available from the MSC. Call Marie Colbert at
905-525-9140 ext. 22496 or write Marie Colbert, Dept. of Pathology,
McMaster Univ. 1200 Main St., Hamilton, Ont. L8N 3Z5. I believe the
Directory may also be available on disc.

Carolyn Emerson, Editor, MSC Bulletin,
Dept. of Biology,
Memorial Univ.
St. John's, Nfld, Canada A1B 3X9.




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Mon, 24 Oct 1994 9:30:38 -0500 (CDT)
Subject: JEOL JSM-25 available for donation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






I have seen this microscope, and occasionally buy equipment
from Driftwood Associates. I have no other connection with
this transaction.
Julian Smith III
Biology
Winthrop University
Rock Hill, SC
VOX 803 323-2111




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Mon, 24 Oct 1994 10:07:37 -0500 (CDT)
Subject: Instrument Lists/International

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


G'day Subscribers...

One forum that might be used for Instrument Lists is the
"Micrsocopy & Microanalysis" World Wide Web server that I
also maintain here at ANL. On it is information about
some of the ANL Microscopy projects, but also sections
dedicated to some of the microscopy/microanalysis societies
(MSA, MAS, RMS, MSC, ASEM) I would consider adding updated lists
to this site or to the Anonymous FTP site here which services
the Microscopy & Microanalysis Software Libraries.

The important thing to remember is that such a listing
would only be a subset of any master lists such as Sandy Silver's
database, or comparable ones elsewhere in the world.

Can I suggest that those who maintain regional instrument
listings contact me off-line. We can then try to set up a "short-form"
which could be uploaded to the WWW site (http:/www.amc.anl.gov) or
the FTP site (www.amc.anl.gov) This would be a start, and would
complement NOT REPLACE each regional listing which would
be more comprehensive. Information on how to obtain the complete
local listing would, of course, be provided.

Again if your a organizer of such list, contact me at

Zaluzec-at-AAEM.AMC.ANL.GOV


When we have something that is on-line, I'll post the
relevent information to the Microscopy Listserver..

Cheers.. Nestor




From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Mon, 24 Oct 1994 10:49:21 -0500
Subject: Wanted: SEM Camera

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Microscopists,

I am presently using a Steinheil Optronic Oscillophot M4, which is "wearing out"
after 16 years of use. I am looking for a new or used (in good working
condition) Polaroid CU-5 camera to mount on my Philips 500X SEM.

Is the CU-5 a good choice, or what other kinds of cameras are people using to
record 4x5 inch pictures? My Steinheil has f/stop, enlargement ratio and
focusing controls which I like. However, it does not quite cover my 4x5 inch
photomonitor screen so there is some loss of detail right in the corners of the
pictures.

Any assistance will be greatly appreciated. Thanks.

--

Gib Ahlstrand
Electron Optical Facility
University of Minnesota
Dept. Plant Pathology
495 Borlaug Hall
St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX
giba-at-puccini.crl.umn.edu






From: MicroToday-at-aol.com
Date: Mon, 24 Oct 1994 15:54:13 -0400
Subject: Microscopy Today

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Goodday!
In response to my recent invitation for no cost subscriptions to our
newsletter, I am overwelmed with questions. With apologies for the "clutter"
to others I would like to address all questions:
1) There is no email version. The 10 yearly issues are available only
through the mail so I will need full and proper mailing addresses.
2) There is currently no charge for subscriptions. We may have to charge a
modest fee in the future for international subscriptions.
3) The objective of the newsletter, perhaps unlike any other, is to publish
material and articles of interest to the working microscopist. It is not a
peer review journal and, while we need ad income to publish it free, it is
not an advertising catelog.
4) A number of members of this listserver do contribute to the effort and
their help is much appreciated. We would greatly appreciate
material/articles from others on the list. If interested, please contact me
direct and I will provide other guidelines.
5) A new feature in the pub is a series of "Tricks of the Trade" notes - on
any microscopy technique and of any length. Authors of each accepted "trick"
will be entered in a drawing at next years MSA Conference for a 1 oz gold
coin. Do two and you should have between a 1 and 25 to 1 and 50 chance of
wining the some $500 worth coin.
Regards,
Don Grimes, Editor - Microscopy Today




From: Liang, Long :      LLIANG-at-is.Arco.COM
Date: 24 Oct 1994 16:21:16 CST
Subject: Coater-- Vendor's address needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to order some accessories for my EMSCOPE SC650 sputter
coater. Does anyone know the address and telephone number of their
vendors in the USA ? Thanks in advance.


Long Liang
Electron Microprobe and SEM Lab
ARCO Exploration and Production Technology
2300 West Plano Parkway
Plano, TX 75075
E-mail : LLIANG-at-is.Arco.COM






From: tivol-at-tethys.ph.albany.edu
Date: Mon, 24 Oct 1994 17:51:41 EDT
Subject: SEM of powder slurries

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Robin,
I have just started to look at clay mineral particles on the HVEM, a
TEM rather than an SEM. The colleague investigating clay from the bottom of
the Hudson river collects a sample of the sediment, washes it and settles it.
During the settling, she collects fractions at various times; these are label-
led for their nominal sizes-- } 5mu, 2mu, {2mu and { {2mu. I had difficulties
getting good dispersal on a carbon-coated mesh grid at first, but I found that
if I put a drop of poly-l-lysine on the grid and let it dry, I could then agi-
tate the suspension of clay particles and put a 5 mu-liter drop on the grid.
This gave me very good dispersion of both the largest and the smallest parti-
cles. I must emphasize that this is my first venture with anything like this,
so listen to the real experts if any respond. (I forgot to say that the 5 mu-
liter drop I put on the grid was just air dried.) I am aware that clay mine-
rals' structures change when they dry, but we wanted to try the simplest thing
first. Good luck.
Yours,
Bill Tivol




From: Loling Song :      loling-at-ruly46.LeidenUniv.nl
Date: Tue, 25 Oct 1994 08:33:48 +0000 (WET)
Subject: athrancene

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to know if anyone from this group has ever used anthrancene
as a dye or a triplet energy acceptor. If so, I would appreacite any
information on its spectral properties, chemsitry, and literature
on its triplet energy.

Thank you in advance!

Loling

===============================================================================
Loling Song Office tel: +31+71-276198
Department of Cytochemistry and Cytometry, +31+71-276200
Leiden University, fax: +31+71-276180
Wassenaarseweg 72,
2333 AL Leiden,
The Netherlands email: song-at-RULLF2.LeidenUniv.NL
===============================================================================





From: Loling Song :      loling-at-ruly46.LeidenUniv.nl
Date: Tue, 25 Oct 1994 09:08:23 +0000 (WET)
Subject: anthracene

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to know if anyone from this group has ever used anthrancene
as a dye or a triplet energy acceptor. If so, I would appreacite any
information on its spectral properties, chemsitry, and literature
on its triplet energy.

Thank you in advance!

Loling

===============================================================================
Loling Song Office tel: +31+71-276198
Department of Cytochemistry and Cytometry, +31+71-276200
Leiden University, fax: +31+71-276180
Wassenaarseweg 72,
2333 AL Leiden,
The Netherlands email: song-at-RULLF2.LeidenUniv.NL
===============================================================================





From: IAN HALLETT :      ihallett-at-marccri.marc.cri.nz
Date: Tue, 25 Oct 1994 22:25:58 GMT+1200
Subject: Instrument list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think Nestor's idea has a lot of merit. Like most other countries
New Zealand has a local directory organised by the newsletter of the
NZ Society for Electron Microscopy which could probably be
incorporated in some international list

Access to a wider range of instrument users would be very useful as
frequently we find we have no one we can ask locally about new or
unusual equipment.
Ian Hallett
HortResearch
Mt Albert Research Centre
Private Bag 92 169
Auckland, New Zealand
Fax 64-9-815 4201
Telephone 64-9-849 3660




From: POSHEL-at-wpo.it.luc.edu
Date: Tue, 25 Oct 1994 08:36:37 -0500
Subject: For R. Crang at U Illinois

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


R. Crang;
I've lost your email address--apologies to the list.
The Neutral Red idea I sent you awhile back for looking at wood cells
has been stewing in the back of my mind, and it's lately bubbled up that
Vaughn was using not Neutral Red, but Rose Bengal. This should also be
in Conn's Biological Stains.
Phil Oshel
poshel-at-luc.edu





From: EMLAB-at-opus.mco.edu
Date: Tue, 25 Oct 1994 13:59:00 -0400 (EDT)
Subject: sections falling off grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {Chameleon.941025091548.tonygr-at-emlab.mit.edu}


Help,

I have been trying to immunogold label some kidney tissue embedded in LR White,
but after three trys, (30 grids) I have only had three sections remain on the
grids, (one section on three grids). I am using gold hex mesh grids,
immersion of grids in 30ul drops of reagents. The grids were precleaned with
detergent, rinsed with dH2O, rinsed with ETOH, and finally with acetone. Why
are the sections falling off??? Any and all comments and suggestions are
welcome.

Thanks
Ed Calomeni




From: XiaoGuang Ning :      ningx-at-mcmail.cis.mcmaster.ca
Date: Tue, 25 Oct 1994 17:02:52 +0059 (EDT)
Subject: help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings!

I have read one paper in certain journal about dislocations in Si3N4,
about which I forget the names of the journal and authors. In that paper,
the Burges vectors of several kinds of dislocations in Si3N4 were
determined by TEM. Could you please tell me related information if you
wrote or have read the paper. In the paper, it was also mentioned that
the dislocations in si3N4 can move only at very high temperature (1700C?).

Best Regards

XiaoGuang NING





From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Wed, 26 Oct 1994 13:05:51 +1300
Subject: Re: sections falling off grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Help,
} Why are the sections falling off??? Any and all comments and suggestions are
} welcome.
}
} Thanks
} Ed Calomeni
}
Ed,
I am informed by that your grids should be formvar coated (prior to putting
the sections on), this should help the sections stick and apparently makes
no difference to the labelling efficiency (as you might expect given that
only one side of the section is exposed to the reagents).


Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: John M Hudak :      hudakjm-at-mcmail.cis.mcmaster.ca
Date: Tue, 25 Oct 1994 14:51:06 +0059 (EDT)
Subject: SEM of powder slurries.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




We quite frequently have to do SEM of many different kinds of
powders. The problem with putting a drop of suspended particles on a stub
is that surface tension draws the particles into clumps as the liquid
dries. This is the technique that I use:
1. Make a very dilute suspension of the particles in methanol.
You may have to determine the concentration by trial and error
but it is very low.
2. Set up a vacuum filtration system using 10mm dia. Nuclepore
filter membranes. I use filters with a pore size of 0.1um.
3. Vacuum filter a few drops of the suspension through the
membrane.
4. Mount the filter on a stub and coat.

By vacuum filtering, the liquid leaves the particles almost
immediately so that surface tension doesn't have a chance to form
clumps. Nuclepore is the best filter to use because it is a sieve with
small holes on a flat background which makes it easier to see the particles.

____________________________________________________________________
John Hudak hudakjm-at-mcmaster.ca
I.M.R. - Electron Optics
McMaster University, Hamilton, Ontario, Canada






From: {rgriffin-at-eng.uab.edu}:ddn:wpafb
Date: 10-24-94 4:40pm
Subject: SEM of powder slurries

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410251526.AA25717-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: SEM of powder slurries
Orig-Author: {"Griffin, Robin" {rgriffin-at-eng.uab.edu} }:ddn:wpafb
-----------------------------------------------------------

Help!

We would like to characterize ceramic powders that are in an aqueous
solution. We need to know if the powders are individual powders or
agglomerates. Does anyone have any suggestion on SEM sample prep? Is
there a drying technique that leaves the particles well dispersed?

Also, what is the typical TEM sample prep techniques used for ceramic
powders?












From: {rgriffin-at-eng.uab.edu}:ddn:wpafb
Date: 10-24-94 4:40pm
Subject: SEM of powder slurries

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410251656.AA26183-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: SEM of powder slurries
Orig-Author: {"Griffin, Robin" {rgriffin-at-eng.uab.edu} }:ddn:wpafb
-----------------------------------------------------------

Help!

We would like to characterize ceramic powders that are in an aqueous
solution. We need to know if the powders are individual powders or
agglomerates. Does anyone have any suggestion on SEM sample prep? Is
there a drying technique that leaves the particles well dispersed?

Also, what is the typical TEM sample prep techniques used for ceramic
powders?












From: {rgriffin-at-eng.uab.edu}:ddn:wpafb
Date: 10-24-94 4:40pm
Subject: SEM of powder slurries

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410251948.AA27076-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: SEM of powder slurries
Orig-Author: {"Griffin, Robin" {rgriffin-at-eng.uab.edu} }:ddn:wpafb
-----------------------------------------------------------

Help!

We would like to characterize ceramic powders that are in an aqueous
solution. We need to know if the powders are individual powders or
agglomerates. Does anyone have any suggestion on SEM sample prep? Is
there a drying technique that leaves the particles well dispersed?

Also, what is the typical TEM sample prep techniques used for ceramic
powders?












From: /S=JOST/OU=BIO/-at-IO-WARNEMUENDE.D400.de
Date: Wed, 26 Oct 1994 03:35:18 -0500
Subject: Help removing mucopolysaccharide films of cyanobacteria for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists
we have a problem with SEM of cyanobacteria filtered onto Nucleopore membrane filter.
Perhaps somebody has a good suggestion for gently removing the mucopolysaccharide (?)
film around the filamentous cyanobacteria to get good pictures of the cell surfaces.
Any suggestions will be very much appreciated!

Guenter Jost
Institute for Baltic Sea Research
D 18119 Warnemuende
Germany
e-Mail Jost-at-bio.io-warnemuende.d400.de




From: rms-at-vax.ox.ac.uk
Date: Wed, 26 Oct 1994 10:36:22 +0000
Subject: RMS Special Interest Groups

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk
JTERLET-at-CEMMA.ADELAIDE.EDU.AU, microtoday-at-aol.com, ROSS.MACKENZIE-at-OX.AC.UK,
MICROSCOPY-at-AAEM.AMC.ANL.GOV
Message-ID: {00986845.E2B5CCDC.14-at-vax.ox.ac.uk}

RMS SPECIAL INTEREST GROUPS IN DIGITAL IMAGING & 3D MICROSCOPY

*********************************************************************

The Royal Microscopical Society has set up two new special interest groups in
Digital Imaging and Three-Dimensional Microscopy. The purpose of the groups is
to organize meetings, workshops, seminars and courses with the support and
backup of the RMS. It is anticipated that, unlike the Sections of the RMS,
these special interest groups will 'wax and wane', or form new permanent
Sections, as the occasion demands.

If you would like to be added to the mailing lists for these groups, please
send your name and address (and email address) to the Royal Microscopical
Society, 37/38 St Clements, Oxford OX4 1AJ, UK. Email RMS-at-VAX.OX.AC.UK. Please
indicate which group you would like to receive information about.

If you have any suggestions/input for these groups, please contact the group
convenors.

DIGITAL IMAGING GROUP CONVENOR: Dr Dominique Jeulin, Centre de Geostatistique,
ENSMP, 35 rue St Honore, F-77300 Fontainebleau, France. Email:
JEULIN-at-CG.ENSMP.FR

3-D MICROSCOPY GROUP CONVENOR: Dr Alan Entwistle, Ludwig Institute for Cancer
Research, Courtauld Building, 91 Riding House Street, London W1P 8BT, UK.

*****************************************************************************





From: EMLAB-at-opus.mco.edu
Date: Wed, 26 Oct 1994 09:05:24 -0400 (EDT)
Subject: Re: LaB6 FIX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Paul,

We have had several LaB6 cathodes which have not quite stopped emmitting but
were close. Check out the mount, these are usually made of tungston or carbon
both of which will decay faster than the LaB6 crystal will. We have tried
sending them back to the maker for warranty replacement but to no avail.
Best Bet--buy a new one or switch back to tungston filaments.

Ed Calomeni




From: Yi Feng :      FENG-at-coefac.engr.wisc.edu
Date: Wed, 26 Oct 1994 09:38:35 CDT
Subject: unsubscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

unsuscription




From: murphy-at-ms.sjdccd.cc.ca.us (Murphy, Judy)
Date: Wed, 26 Oct 1994 09:03:11 PST
Subject: RE: Sections falling off

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {1994Oct26.090311.1567169635-at-ms.sjdccd.cc.ca.us}
To: microscopy-at-aaem.amc.anl.gov (MSA list)

To eliminate the problem of the sections falling off, for years we have used
"grid glue". Take a 1 inch piece of double sticky tape and put it in 20
ml. of ethylene dichloride. Wait a few min. until the adhesive comes off
and then remove the piece of tape. In a closed bottle this stays good for
months. We dip the cleaned grids in the grid glue, let dry and use the
grids for picking up sections. We never use formvar coated grids for
sections, but always do use grid glue. Prior to use we clean all of our
grids in a small amount of acetone in an ultrasonic cleaner for a minute or
so. There is some sort of residue on the grids when they are new which is
then removed. We have used all EM supply houses, and found cleaning
necessary before picking up sections. Actually we clean all of our grids
for any use, i.e. whole mounts, or sections. I believe I published this
little trick some years ago, 1982 in an SEM paper on Mounting materials
where we used the grid glue for a different purpose. Hope it works for you
also. I should add that we use a variety of resins, so don't think it is
resin dependent necessarily. We have used LR White, Embed, the old Epon,
Spurrs, Araldite, etc.

Good Luck
Judy M.

Judy Murphy, PhD
Dept. of Microscopy
San Joaquin Delta College
5151 Pacific Ave
Stockton, CA 95207

Phone: 209/474-5284
FAX: 209/474-5649
e-mail: murphy.ms.sjdccd.cc.ca.us







From: Michael Rock :      merock-at-u.washington.edu
Date: Wed, 26 Oct 1994 09:20:51 -0700 (PDT)
Subject: Re: sections falling off grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


for sections falling off grids, try dipping your (clean) grids in Butvar
a 0.10% solution (in chloroform),
then allow to dry on a clean piece of glass or a fiberless/ashless piece
of filter paper. store in a clean dry place.
both surfaces of the section will be available for immuno reagents, since
the butvar is only on the grid bars





From: lmiller-at-ux1.cso.uiuc.edu
Date: Wed, 26 Oct 1994 11:24:56 -0500
Subject: Sections falling off

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,
I've worked with LR White and immunogold, and have not often
had this problem even with days of incubations.

1. The grids, I use Ni, 200 hex. I use only alcohol to clean
them.
In fact, I store the grids in 100% EtOH. I suspect that the
acetone
could be a problem if it is not washed off very very well. We've
found that acetone leaves a residue on our grids, that is why we
use only ethanol.*** Also be sure , just prior to cutting, that
you wash the grids off with about a ml of water, then dry right
up to a 75 watt bulb. We do this to be sure the EtOh is gone,
but also because of "static" problems we have in the lab.

2. If the antigen site is not too heat sensitive, imediately
after wicking the grids to " just damp", I hold the back
side of the grid up 2mm away from a bare, 75 watt
lamp. This is done only 5 seconds longer than all
visable moisture is gone and the grid looks dry.
** This has kept our sections on and flat, even with
" jet" washing from a water bottle.

3. How thick are your sections, very thick sections will
fall off of the grid more easily. I cut LR White at
65-80nm.
If this is hard to cut, go back to the embedding to improve
the
block.
**** when I finish polymerizing the LR White, I put it
into a desicater jar, and put it under vacuum for 3-4 days
before sectioning. Storage is also under vacuum.
For some reason, this seems to help a lot.
Good Luck!
Lou Ann
---------------------
Lou Ann Miller
U of Illinois
College of Veterinary Medicine
217-244-1566
lmiller-at-ux1.cso.uiuc.edu







From: Joyce Craig :      bafpjec-at-uxa.ecn.bgu.edu
Date: Tuesday, November 15, 1994
Subject: scanning probe free workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Midwest Society of Electron Microscopists (MSEM) presents a Free*
workshop: Scanning Probe Microscopy of Materials and Phenomena

Place: Northwestern University, Norris Center Room 2A
8:30 Registration
9:00 - 12:00 Speakers:
Dr. Z. Zhang, Universiity of Wisconsin-Madison
Mr. Ian Smith, Park Scientific Instruments
Dr. Jerry Zajac of Amoco
Dr. Carla Alves from Topometrix
Dr. John B. Ketterson, Northwestern University
11:45 - 1:00 Lunch (Norris Center Cafeteria)
1:00 - 5:00
Panel Discussion: "Where to from Here"
Vendor displays and demonstrations
Student Poster Presentations
Facility tour

For reservations and further information call:
Joyce Craig or Jeffrey Schmelz
E-Mail BAFPJEC-at-UXA.ECN.BGU.EDU
fAX 312-995-3759
pHONE 312 995-3800
FOR INFORMATION ABOUT THE STUDENT PRESNTATIONS CONTACT MEETING ORGANIZER:
VINAYAK P. DRAVID
e-MAIL VPDRAVID-at-CASBAH.ACNS.NWU.EDU
708 467-1363
708 467 7798
fax 708 491-7820

*The workshop is free to MSEM members. Ask for an application.
Membership is only $10 per year, $5 for students.




From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Wed, 26 Oct 1994 16:16:23 -0400
Subject: sections falling off

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To keep sections on the grid, I usually clean my grids just before
picking up the sections. I use a squirt bottle with 10% acetic acid
followed by a distilled water rinse and then an acetone rinse. I do
this by gripping the grid in my tweezers and squirting the grid with the
different solutions and then putting the grid down on a piece of #50 Whatman
filter paper. I never have a problem with the sections falling off. The
grids I use mostly are the standard copper grids. I use the same
procedure for nickel grids except after the acetone wash I do a second
distilled water rinse.
Hope this helps.
Phil
8-{)




From: kayton-at-ohsu.edu (Robert Kayton,MAC,CROET)
Date: Wed Oct 26 14:54:16 PDT 1994
Subject: PNEMS Fall 1994 meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0r0GIS-0007KIC-at-stjohns.ohsu.edu}
Message-Version: 2
} To: Microscopy-at-AAEM.AMC.ANL.GOV


Mark your Calendars



Portland area's Pacific Northwest Electron Microscopy Societies mini-meeting
will be held on Tuesday, November 1, 1994. The meeting agenda includes a
buffet at Shang Hai Nobel House restaurant a short PNEMS business meeting and
a presentation by OPTIX Inc.. The restaurant is located at John's Landing
5331 S. W. MacAdam Ave, just south of downtown Portland.

The business meetings topics include the spring '95' PNEMS meeting, MSA 1999
national meeting, and miscellaneous items. The presentation will cover
imaging topics from archiving large digital files to 3-dimensional
reconstruction. Buffet begins at 6:00 PM and the presentation will conclude
around 8:00 PM.

Since the Society will be providing the buffet please RSVP to Bob Kayton
(503)494-2504 by October 27. Please contact Bob Kayton if you have any
questions. Any interested microscopists are encouraged to attend.

See you there!




From: kayton-at-ohsu.edu (Robert Kayton,MAC,CROET)
Date: Wed Oct 26 16:02:59 PDT 1994
Subject: PNEMS Fall 1994 meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0r0HMx-0007KEC-at-stjohns.ohsu.edu}
Message-Version: 2
} To: microscopy-at-aaem.amc.anl.gov


Mark your Calendars



Portland area's Pacific Northwest Electron Microscopy Societies mini-meeting
will be held on Tuesday, November 1, 1994. The meeting agenda includes a
buffet at Shang Hai Nobel House restaurant a short PNEMS business meeting and
a presentation by OPTIX Inc.. The restaurant is located at John's Landing
5331 S. W. MacAdam Ave, just south of downtown Portland.

The business meetings topics include the spring '95' PNEMS meeting, MSA 1999
national meeting, and miscellaneous items. The presentation will cover
imaging topics from archiving large digital files to 3-dimensional
reconstruction. Buffet begins at 6:00 PM and the presentation will conclude
around 8:00 PM.

Since the Society will be providing the buffet please RSVP to Bob Kayton
(503)494-2504 by October 27. Please contact Bob Kayton if you have any
questions. Any interested microscopists are encouraged to attend.

See you there!




From: Daniel Henne :      henne-at-sfu.ca
Date: Wed, 26 Oct 1994 16:19:29 -0700 (PDT)
Subject: EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Two questions concerning X-ray analysis from a desperate Masters
student.
1. Has anybody worked with the ZAF program QUAD2 developed by
Farthing et al and presented at the Int. Congr. X-ray Optics and
Microanalysis, Manchester, 1992.
2. How and what is an Electron Probe Microanalysis. We have our EDX
on an SEM and I understand how that set-up does it's compositional
analysis but I don't understand how standards are used. Let me
explain that I understand most of the equations in the ZAF analysis
and I have read the standard texts but I think I'm missing something
on the operations.
Replying to me directly can be done at henne-at-sfu.ca
Thanks in advance.
Cheers.
Dan Henne
Simon Fraser University
Vancouver, Canada





From: kayton-at-ohsu.edu (Robert Kayton,MAC,CROET)
Date: Wed Oct 26 16:26:15 PDT 1994
Subject: PNEMS Fall 1994 meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0r0HjT-0007KFC-at-stjohns.ohsu.edu}
Message-Version: 2
} To: microscopy-at-aaem.amc.anl.gov

MEETING

Hi,

The Portland area's Pacific Northwest Electron Microscopy Society is having
its first mini-meeting on Tuesday November1, 1994. The meeting agenda
includes a bufet at Shanghai Nobel House restaurant a short PNEMS business
meeting and a presentation by OPTIX Inc. The restaurant is located at John's
Landing, 5331 S.W. MacAdam Ave., just south of downtown Portland.

The business meetings' topics include the spring '95' PNEMS meeting, MAS 1999
national meeting, and miscellaneous items. The presentation will cover
imaging topics from archiving large digital files to 3-dimensional
reconstruction. Buffet begins at 6:00 PM and the presentation will conclude
around 8:00.

Since the Society will be providing the buffet please RSVP to Bob Kayton
(503)494-2504 by Nov. 1. Please contact Bob Kayton if you have any questions.

E-mail address kayton-at-ohsu.edu




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Wed, 26 Oct 1994 20:38:47 -0500 (CDT)
Subject: duplicates

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


G'day subscribers,


Just to let you all know I have noticed duplicate and sometimes
triplicate messages on the net. Looking at the headers they appear
to be coming from the originators, the majority of the time. There
are some that are mirrored by the system if the computer crashes
during a session, but that has become less frequent with the
new system. All I can say is sorry :-( for the traffic.
If anyone is not sure about how to send a message
touch base with me off-line. Remember the mail queue sometimes
gets very long and if you post a message it may not appear
on the net for several hours, or sometimes not until the next
day. Just be patient, it will get there!


Just for the record the Microscopy Listserver has just passed
it's first anniversity/birthday. We started operation on
Oct 1, 1993 and the system has long ago passed the point of it's
1,000,000th Email message delivery. To date we've
delivered over 2500 postings/messages to over 1500 subscribers
(plus an unknown number of readers via the SciTechnqiues Microscopy
Newsgroup) 1M is conservative since by simple math 2.5Kx 1.5K = 3.75M,
but not every subscriber has seen every message
(except, of course, for me). Not bad for our first year of operation.
I don't actually know who got the 1Mth and I doubt if I could find out.
In any case just bear with the glitches and things will work out in
the long run.

Cheers

Nestor







From: david.rayns-at-stonebow.otago.ac.nz (David Rayns)
Date: Thu, 27 Oct 1994 15:57:06 +1200
Subject: Fluorescent microtubules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199410270253.AA11418-at-arwen.otago.ac.nz}
X-Sender: st004084-at-brandywine.otago.ac.nz
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Does anyone know if there is an 'off the shelf' fluorescent drug or
compound analogous to FITC-phalloidin, but which will react with
microtubules? Thanks for your help. David.






From: montpetitd-at-EM.AGR.CA
Date: Thu, 27 Oct 1994 08:46:39 -0400
Subject: sections falling off grid response

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {seaf6895.059-at-EM.AGR.CA}
X-Mailer: WordPerfect Office 4.0

hi,
I would coat my grids with formvar to prevent sections
from falling out.

Diane Montpetit
Food research center
St-Hyacinthe, quebec, Canada





From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Thu, 27 Oct 1994 09:08:40 -0400
Subject: LR-White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anybody out there do flat embedding of tissue embedded in LR-White?
I know the embedding mold has to be covered during the curing to keep
the air out but do they make a special mold for doing this? All I have
are the standard embedding molds for flat embedding. Can these be used
somehow?
Thanks in advance,
Phil
8-{)




From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Thu, 27 Oct 1994 09:52:11 -0400 (EDT)
Subject: RE: LR-White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Thu, 27 Oct 1994 09:08:40 -0400,
rutledge phil {prutle1-at-gl.umbc.edu} writes:

} Does anybody out there do flat embedding of tissue embedded in LR-White?
} I know the embedding mold has to be covered during the curing to keep
} the air out but do they make a special mold for doing this? All I have
} are the standard embedding molds for flat embedding. Can these be used
} somehow?
} Thanks in advance,
} Phil
} 8-{)
}
*************

Yes, Ted Pella (USA 1-800-637-3526; Canada 1-800-243-7765) supplies teflon
molds (cat# 10506) and ACLAR film (cat# 10502) that provide an oxygen-free
environment for the polymerixation of LR White & Lowicryls.

*****************************************************
M.V. Parthasarathy
Professor of Plant Biology; Director, EM Facility
Section of Plant Biology, 228 Plant Science Building
Cornell University, Ithaca, NY 14853. USA
Tel: 607-255-1734; Fax: 607-255-5407
E-Mail: mvp2-at-cornell.edu
******************************************************
e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-
e- e-
e- This message was sent using 100% recycled electrons e-
e- e-
e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-e-




From: W.L. Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Thu, 27 Oct 1994 09:45:42 EST
Subject: LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Phil,
I have been using silicone embedding molds for years for embedding LR
White. After filling the molds, I place it into an airtight container
which is flushed with an inert gas (dry nitrogen, argon, freon, all work).
The cover is installed and the container placed in the embedding oven. I
use a 1 pint paint can with a hole in the top for introducing the gas.
Once flushed, the hole is covered with a piece of tape. This system works
just fine.

-=W.L. Steffens=-
College of Veterinary Medicine
University of Georgia




From: Greg Erdos ICBR EM Core Lab Univers :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Thu, 27 Oct 1994 09:51:39 -0500 (EST)
Subject: Fw: Re: LR-White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


=From: TOWER::GWERDOS "Greg Erdos ICBR EM Core Lab Univers"
=To: IN%"prutle1-at-gl.umbc.edu"
=CC: GWERDOS
=Subj: Re: LR-White
=
=} From: IN%"prutle1-at-gl.umbc.edu" "rutledge phil"
=} Subj: LR-White
=}
=} Return-path: {prutle1-at-gl.umbc.edu}
=} Received: from AAEM.AMC.ANL.GOV by gnv.ifas.ufl.edu (PMDF V4.3-10 #3240)
=} id {01HIRLP5TFJ48XUR11-at-gnv.ifas.ufl.edu} ; Thu, 27 Oct 1994 09:37:58 -0500 (EST)
=} Date: Thu, 27 Oct 1994 09:08:40 -0400
=} From: rutledge phil {prutle1-at-gl.umbc.edu}
=} Subject: LR-White
=} X-Sender: prutle1-at-umbc8.umbc.edu
=} To: microscopy {microscopy-at-aaem.amc.anl.gov}
=} Message-id: {Pine.SGI.3.90.941027090342.4522A-100000-at-umbc8.umbc.edu}
=} MIME-version: 1.0
=} Content-type: TEXT/PLAIN; charset=US-ASCII
=} Content-transfer-encoding: 7BIT
=}
=} Does anybody out there do flat embedding of tissue embedded in LR-White?
=} I know the embedding mold has to be covered during the curing to keep
=} the air out but do they make a special mold for doing this? All I have
=} are the standard embedding molds for flat embedding. Can these be used
=} somehow?
=} Thanks in advance,
=} Phil
=} 8-{)
=#######################################################
=Phil,
= For LR White etc. where polymerization is oxygen sensitive. I punch out
=circles of unexposed but cleared EM film with a paper punch. I put these under
=resin in a polypropylene microfuge tube. They settle part way down the taper
=and form a flat platform for tissue. The tube can be nealy filled with resin,
=closed and successfully polymerized. Then you cut the tube down the side,
=remove the solid plastic and break it at the film exposing your tissue at a
=flat surface ready for sectioning.
= Send an address or a FAX number and I can provide more details and
=references.
=**********************************************************
=* Greg Erdos ** *
=* Director, ICBR EMCL ** Phone 904-392-1295 *
=* 218 Carr Hall ** FAX 904-846-0251 *
=* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
=* Gainesville, FL 32611 ** *
=****************************************************************
**********************************************************
* Greg Erdos ** *
* Director, ICBR EMCL ** Phone 904-392-1295 *
* 218 Carr Hall ** FAX 904-846-0251 *
* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
* Gainesville, FL 32611 ** *
****************************************************************




From: gkennedy-at-ucsd.edu
Date: Thu, 27 Oct 1994 08:31:36 -0800
Subject: LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199410271528.IAA05400-at-ucsd.edu}
X-Sender: gkennedy-at-popmail.ucsd.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
Content-Transfer-Encoding: quoted-printable

I flat-embed regularly in LR White, between 2 sheets of Aklar film from Ted=
Pella: I infiltrate my sections (60-80 microns, vibratomed) as usual, them=
immerse them in a small puddle of something like araldite or eponate, etc.=
on the bottom piece of Aklar, then place a second sheet of Aklar on top. =
Bake as usual. Incidentally, this is NOT the material that I had trouble=
with in the beam--that stuff was actually a regular tissue block. I=
finally decided, thanks to the many kind responses I received, that my LR=
White is just too old. Grace






From: Paulette Brunner :      pbrunner-at-u.washington.edu
Date: Thu, 27 Oct 1994 10:33:49 -0700 (PDT)
Subject: recordable CD-ROM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We would like to use recordable CD-ROMs for archiving images from a Gatan
system we will be purchasing. I would be interested in hearing from
anyone using CD-ROMs for storage.

I just saw an ad for a Ricoh CD-ROM unit for approx.$3000.00. It is both
multi-session and multi-platform. Has anyone used this unit?

Please reply to pbrunner-at-u.washington.edu

Thanks,

Paulette Brunner




From: RRATLENG1!RRATLENGPO!ALLSMAN-at-rratleng.attmail.com
Date: Thu, 27 Oct 1994 15:11:00 +0000
Subject: SUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html








From: RRATLENG1!RRATLENGPO!ALLSMAN-at-rratleng.attmail.com
Date: Thu, 27 Oct 1994 15:16:00 +0000
Subject: SUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To whom it may concern:
We have an old Coates & Welter Cwikscan FE-SEM. While talking to
someone about our instrument it came to light that there was a message
on this bulletin board about another Cwikscan which someone was
trying to find a user for.
I am interested in finding out more information about this SEM. Any help
is appreciated.
Thanks,
Jerry Allsman
Rolls-Royce Inc.
Atlanta, GA





From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Thu, 27 Oct 1994 11:19:07 -0700 (PDT)
Subject: Re: LR-White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer09.u.washington.edu

I've flat embedded items in Historesin using Peel-a-Way paraffin molds
and ;bottle caps that are carefully covered ;with a layer of mineral oil
or melted paraffin. Also handy is to make some non-stick microscope
slides by coating them with silane. Then lay down a 100-300 micron slice
of infiltrated tissue and cover ;with methacrylate. lay another coated
slide on top, being careful to not entrap air bubbles. Now paint the
edges with paraffin and allow to polymerize. Sandwiching between slides
is a standard method in our lab for methacrylate embedding of slices.

Glen MacDonald
Hearing Development Laboratories RL-30
University of Washington
Seattle, WA 98195
(206)543-8360
glenmac-at-u.washington.edu



On Thu, 27 Oct 1994, rutledge phil wrote:

} Does anybody out there do flat embedding of tissue embedded in LR-White?
} I know the embedding mold has to be covered during the curing to keep
} the air out but do they make a special mold for doing this? All I have
} are the standard embedding molds for flat embedding. Can these be used
} somehow?
} Thanks in advance,
} Phil
} 8-{)
}




From: BALTRUS-at-petc.doe.gov
Date: Thu Oct 27 17:18:03 1994
Subject: KEPCO Company

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We currently have an old KEPCO Model NTC 2000 power supply in
need of repair. Attempts to call the company at the number
listed on the manual that came with the supply were fruitless.

Does anyone know if this company is still in business and if so,
how I may contact them. Alternatively, is there a company that
can repair the power supply?

Please direct all responses directly to me at
baltrus-at-orion.petc.doe.gov

Thanks for your help!

John Baltrus
US Dept of Energy/Pittsburgh Energy Technology Center





From: Glenn Poirier :      GLENN_P-at-GEOSCI.Lan.McGill.CA
Date: Thu, 27 Oct 1994 16:57:22 EST5EDT
Subject: Microscope labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199410272058.QAA01139-at-sifon.CC.McGill.CA}

Greetings Microscopists,
I'm looking for info on magmetic shielding of microscope labs.
Specifically, we are building a new materials science building, and
would like to make absolutely certain that magnetic fields are
minimized. We are mostly interested in keeping fields from outside the
labs (from machine shops and the like) from causing problems. Any
thoughts or advice on this matter would be appreciated.

Glenn Poirier
Microprobe Lab
McGill University
Montreal, Qc
514 398 6774




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Fri, 28 Oct 1994 10:12:49 +1300
Subject: Help - root nodules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Has anyone out there got any experiance/ideas re root nodules for TEM?

We have prepared clover root nodules for TEM and have had problems getting
good sections - the problem is that regions in the middle of the sections
of the plant cells fall out.. I'm not sure whether the fixation of the
centers is poor or whether its a resin problem. I have found that Agar 100
works better (has fewer holes) than Spurrs which I initially used. I fix
under a slight vacuum for 2 hours to aid infiltration of the 3% glut 0.2M
cacodylate into the 1-2mm diameter nodules, plus slice a portion off the
nodules to open them up a little but to no avail.I use 1% Oso4 in same
buffer as postfix.
The regions which are intact look good however.
Any ides?

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Thu, 27 Oct 1994 16:08:31 -0500
Subject: Minnesota Microscopy Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


MMS FALL BUFFET DINNER & IMPORTANT BUSINESS MEETING

UNIVERSITY OF MINNESOTA, ST. PAUL CAMPUS STUDENT UNION, CHERRYWOOD ROOM,
THURSDAY, NOVEMBER 10, 1994, 5:00 - 9:00 PM

SPEAKER: JAN HINSCH, LEICA INC.

TOPIC: "BUSMAN'S HOLIDAY"

MMS will hold its sixth annual Fall Buffet Dinner on November 10, 1994, at the
University of Minnesota's Cherrywood Room located on the 2nd floor of the St.
Paul Campus Student Union, 2011 Buford Avenue, St. Paul Campus( NOTE: This
year's Buffet Dinner is NOT at the Campus Club as in previous years). We hope to
provide a pleasant evening during which microscopists will be moved to renew or
begin their membership in MMS, MSA and MAS. A wine, cider and cheese social
from 5:00-6:00 in the Cherrywood Room. will kick off the evening. The buffet
dinner follows from 6:00-7:00. The dinner entre will be Halibut Steak with
Chicken Strips (Cacciatore Style), Caesar Salad, Rice Pilaf, Fresh Broccoli
Spears, desserts and beverages. The total cost is more than the $10.00 fee but
MMS is picking up about 40% of the tab as a courtesy to our membership. The
dinner affords an excellent opportunity to meet microscopists from many
disciplines, talk shop, and to have a pleasant time together. The program is
from 7:00-8:00 in the Cherrywood Room. Parking is available behind the Union and
at other St. Paul Campus locations.

Our featured speaker, Jan Hinsch, from 1978 - Present, has been Director of
LEICA's WILD-LEITZ Laboratory for Applied Microscopy in Rockleigh, New Jersey He
is a Fellow, New York Microscopical Society, an Honorary Member and recipient of
the Outstanding Microscopists award from the State Microscopical Society of
Illinois. Jan has had numerous photomicrographs and articles published in a wide
variety of journals - ie. "Industrial Microscopy", 1979 Industrial Research &
Development; "Critical Focusing in Low Power Photomicrography", 1979, American
Laboratory; "Essentials of Light Microscopy and Photomicrography", 1983,
Pathology Annual; "Refined Approaches to Microscopical Light Management", 1988,
The Microscope.
Here is Jan's description of his talk, "Busman's Holiday": "The vacationing
microscopist faces a dilemma. If this is to be a time to conquer the unknown
should not one leave the microscope and any thought of it at home? And yet,
this is also a time of opportunity to collect samples of many kinds and,
inspired by the natural phenomena around, to meditate on the forms and
illustrations of the action of light and its significance to us microccopists.
This is the account of someone trying to have it both ways."
We will hold a short business meeting just before the talk. The Buffet Dinner is
$10.00 per person payable at the door. (non-member $20.00, includes new
membership), $7.00 for current students(non-member student $12.00, includes new
membership). In order for us to provide an accurate head count to the Cherrywood
Room, please make a phone reservation by calling Mike or Ev.

Mike Coscio(612-569-1331, E-MAIL: mike.coscio-at-medtronic.com)
Ev Osten(612-736-0104, E-MAIL: efosten-at-mmm.com) - Buffet Coordinators


--

Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu






From: mrdunlap-at-ucdavis.edu (Michael Dunlap)
Date: Thu, 27 Oct 1994 18:19:22 -0700
Subject: Root Nodules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a had similar problems. We found that air bubbles in the
root nodules, lung, ... prevented the fixation through embedding to work
properly. Try pulling a vacuum on the tissue in the fix after cutting a
small hole in one end. If the tissue floats, clip a staple or small paper
clip to one end to act as a weight. On more difficult samples, we need to
repeat the vacuum a second time when in 100% ETOH (because of the surface
tension in H2O) to get rid of smaller bubbles.
Good luck
Mike

---------------------------------------------------------------------------
| Michael Dunlap | lab (916) 752-0284 |
| Facility For Advanced Instrumentation | fax (510) 422-2282 |
| University of California | mrdunlap-at-ucdavis.edu |
| Davis CA, 95616 | |
===========================================================================






From: anne-at-emu.su.oz.au (Anne Simpson Gomes)
Date: Fri, 28 Oct 1994 11:48:46 +1000
Subject: Re: LR White flat embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410280147.AA08661-at-ELECTRON.emu.su.OZ.AU}
X-Sender: anne-at-electron.emu.su.oz.au (Unverified)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Well I have seen some other solutions to your flat embedding woes and
thought I might as well throw in my 2 cents worth too. At the very least
its another option to try.

Look up my technical paper Flat Mold Embedding with LR White and Lowicryl
K4M. Simpson Gomes,A and Simon G.T. J. of EM Technique 13:266-267 1989.

Note: this journal is now called J. of Microscopy Reseach and Technique.

Have fun, looks like you have several good ideas to work with.

Anne

Anne Simpson Gomes

EM Unit, F09 "How's it going Eh?!!!".......
Univ of Sydney from
NSW 2006 Australia The Compact Canuck!!
Fax: (612) 552 1967







From: anne-at-emu.su.oz.au (Anne Simpson Gomes)
Date: Fri, 28 Oct 1994 12:23:21 +1000
Subject: Re: root nodules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9410280221.AA08730-at-ELECTRON.emu.su.OZ.AU}
X-Sender: anne-at-electron.emu.su.oz.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi

The following info comes from my boss Dr Maret Vesk. She has done extensive
work on clover root nodules.

"Some possible causes:
Air in intracellular spaces = no resin infiltration = holes in sections.
This is very common in plants and may be helped by evacuating roots in
distilled water before fixation. They should sink, not float.
0.2M buffer wiil definitely cause plasmolysis, most workers use 0.025-0.05M
buffer with phosphate rather than the toxic cacodylate being the preferred
buffer.
Use very slow infiltration of resin (Spurr's is perfectly fine), adding
drop wise over a matter of days if necessary.
Is your dehydration long enough (2 x 30min)?
Have had no problems with either clover nodules or wheat paranodules"

Maret (via Anne)


Anne Simpson Gomes

EM Unit, F09 "How's it going Eh?!!!".......
Univ of Sydney from
NSW 2006 Australia The Compact Canuck!!
Fax: (612) 552 1967







From: Dirk Knoesen :      D.Knoesen-at-fys.ruu.nl
Date: Fri, 28 Oct 1994 09:16:39 +0100 (GMT+0100)
Subject: recordable CD-ROM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The EM-unit here at Utrecht Univ, Biology Dept is using a CD-ROM
system for recording images. They are writing it with a Kodak DCD
Writer 200 Plus unit, multisession. I am saving XTEM images with it,
which I read here in my lab on a Panasonic CD-Rom driver. I am using
a 486, 66MHz PC and either the CorelDraw or PhotoShop packages to
analyze the images, primarily for morphological details, and it works
fine.

Dirk Knoesen, Debye Institute, Dept Atomic and Interface Physics,
Uuniversity Utrecht.





From: MikGu-at-mme.liu.se (Mikael Gustafsson)
Date: Fri, 28 Oct 1994 10:51:11 +0100
Subject: FURA-2 imaging.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To Calcium imagers...

I just wonder if anyone knows about immersion oils that transmit 340 nm
light well. Zeiss and Nikon immersion oils attenuates this wavelength
heavily, and the attenuation is dependent of the thickness of the oil layer,
leading to unstable calibrations from time to time and where the focus is in
the sample. I now use glycerol which transmits nice in 340, but this is not
optically optimal because of its lower refractive index.

=============================================
Mikael Gustafsson MD, PhD
Dept Med. Microbiology and
Dept Internal Medicine, Cardiology section
University of Linkoping
SWEDEN

E-Mail: MikGu-at-mme.liu.se
FAX: 046/13/224789
Phone: 046/13/224783
=============================================





From: MikGu-at-mme.liu.se (Mikael Gustafsson)
Date: Fri, 28 Oct 1994 11:16:55 +0100
Subject: FURA-2 imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To Calcium imagers...

I just wonder if anyone knows about immersion oils that transmit 340 nm
light well. Zeiss and Nikon immersion oils attenuates this wavelength
heavily, and the attenuation is dependent of the thickness of the oil layer,
leading to unstable calibrations from time to time and where the focus is in
the sample. I now use glycerol which transmits nice in 340, but this is not
optically optimal because of its lower refractive index.

=============================================
Mikael Gustafsson MD, PhD
Dept Med. Microbiology and
Dept Internal Medicine, Cardiology section
University of Linkoping
SWEDEN

E-Mail: MikGu-at-mme.liu.se
FAX: 046/13/224789
Phone: 046/13/224783
=============================================





From: RRATLENG1!RRATLENGPO!ALLSMAN-at-rratleng.attmail.com
Date: Fri, 28 Oct 1994 07:45:00 +0000
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



send index





From: kleifer-at-cimesg1.epfl.ch (Klaus Leifer)
Date: Fri, 28 Oct 1994 13:46:56 +0000
Subject: Change of e-mail address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: Microscopy Mailing List {microscopy-at-aaem.amc.anl.gov}

Could you please change my e-mail address.
Old address: kleifer-at-i2msg1.epfl.ch
new address:kleifer-at-cimesg1.epfl.ch
thanks
Klaus Leifer

__________________________________________________________________
Klaus Leifer, Ecole Polytechnique Federale de Lausanne (EPFL)
Address: EPFL-I2M, CH-1015 Lausanne, Switzerland
Phone: +41(21)693 48 30 Fax: +41(21)693 44 01
__________________________________________________________________






From: kleifer-at-cimesg1.epfl.ch (Klaus Leifer)
Date: Fri, 28 Oct 1994 15:10:49 +0000
Subject: SAD precision 2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


}
} To:Bob Keller {keller-at-micros.mrd.bldrdoc.gov}
} From:kleifer-at-cimesg1.epfl.ch (Klaus Leifer)
} Subject:SAD precision 2
}
} Hallo Bob,
} thanks for your comment which I received one month ago on the microscopy
} server with respect to precision of evaluation of electron diffraction
} patterns. The system, I'm working on has the following properties with resp=
ect
} to diffraction.
} - Ni/Ti multilayers on Si substrate
} - grain sizes about 3 to 15 nm (in both lateral and orthogonal direction wi=
th
} respect to the interfaces)
} - texture with respect to interfaces (with an angular deviation of +-5=B0)
}
} I need a better precision of SAD to measure strains, absolute average latti=
ce
} plane spacings and their variations.
} Another reason is that the opinion I get from people with whom I discuss is
} that the general limit of SAD is 1%. If I then regard the fact that most of
} the people I know (me included) are still evaluating their diff. pattern wi=
th
} a ruler (perhaps the precision of human eye is not so bad, but numerical fi=
ts
} of the pattern similar to X-ray patterns are not yet very current) I think
} that an enhancement of the 1% precision is still possible.
} This was shown i.e. by Y. Le Page (1992, Microscopy Research and Technique,
} Vol23,No3)
}
} SAD presents the advantage, that I do have the calibration of the Si substr=
ate
} directly on the same diffraction pattern.
} CBED on single grains would be possible with respect to evaluation of discr=
ete
} spots positions(-} I'll try this on a Hitachi HF2000 with 1nm FWHM of the
} probe). The difficulty is that I lose Si calibration on the same
} pattern.Changing from diff. to image mode and back to diff. again gives me
} typical deviations of the position of the Si spots of about 0.5%!!
} The HOLZ lines in only 5nm thick cristallites don't have enough intensity.
}
} I was thinking and calculating to get a possible error for the SAD
} measurements. The probable errors, I found are:
} - distortions of the pattern due to lens spiral and barrel distortions (I t=
ry
} to exclude this
} taking for calibration the Si spots close to the spot I evaluate)
} - precision of the measurement of the peak maximum (well, I'm using a ruler=
)
} - form of the grains together with the inclination angle of the grains with
} respect to the
} beam may change the position of the reflection in between 5e-4 (in =
a
} calculation I
} made taking grains that are only limited in size in beam direction)
} and more than 1%
} (for grains which are limited in size only in one direction
} orthogonal or inclined to
} the beam. I don't observe such some grains but cannot exclude the
} influence of
} stacking faults or twins on the diff. pattern).
} - !! one error about which I have no idea is the following: is there a
} correlation between
} the position of a grain in the SAD aperture and a probable change =
of
} the position of
} its reflections in the diffraction pattern due to higher order
} abberations of the
} objective lens which depend on the position of the object? Do there
} exist
} calculations?
} What do you think about this?
} Best wishes
} Klaus
}
}
} } Klaus,
} }
} } There is some discussion of using SAD for lattice spacing measurements in
} } Reimer's TEM book, chapter 9. Generally, that's not a good approach if yo=
u
} } would like to determine lattice constants with better than a 1 percent err=
or
} } or
} } so. This is a significant error if you are concerned with elastic strain
} } measurements or phase identification. In practice, unless you use a
} } calibration
} } standard, the camera length will not be known accurately and further, ther=
e
} } is
} } also some distortion in the pattern caused by the projector lens.
} }
} } It is not possible to use SAD to determine lattice constants in the direct=
ion
} } of
} } the incident beam. CBED will give you such info. since HOLZ reflections c=
an
} } be
} } excited. Using a STEM unit, it becomes possible to produce CBED spot size=
s
} } around 5 nm dia. fairly routinely, which would allow probing of individual
} } grains in some multilayer systems, presuming you can get single grains
} } through
} } the foil thickness.
} }
} } I recognize from your email address that you are at the same institute as =
Dr.
} } P.
} } Stadelmann. He has written some very nice simulation software that may al=
low
} } you to experiment with how sensitive spot patterns are compared to CBED HO=
LZ
} } patterns for lattice constant measurement.
} }
} } Regards,
} }
} } Bob Keller
} } NIST
}

__________________________________________________________________
Klaus Leifer, Ecole Polytechnique Federale de Lausanne (EPFL)
Address: EPFL-I2M, CH-1015 Lausanne, Switzerland
Phone: +41(21)693 48 30 Fax: +41(21)693 44 01
__________________________________________________________________






From: kleifer-at-cimesg1.epfl.ch (Klaus Leifer)
Date: Fri, 28 Oct 1994 15:17:32 +0000
Subject: new e-mail address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: Microscopy Mailing List {microscopy-at-aaem.amc.anl.gov}

Could you please change my e-mail address.
Old address: kleifer-at-i2msg1.epfl.ch
new address:kleifer-at-cimesg1.epfl.ch
thanks
Klaus Leifer

__________________________________________________________________
Klaus Leifer, Ecole Polytechnique Federale de Lausanne (EPFL)
Address: EPFL-I2M, CH-1015 Lausanne, Switzerland
Phone: +41(21)693 48 30 Fax: +41(21)693 44 01
__________________________________________________________________






From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 28 Oct 1994 11:45:14 U
Subject: RE- MagShields for labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1428777678.56297-at-mse.engin.umich.edu}

Subject: Time:11:28 PM
OFFICE MEMO RE: MagShields for labs Date:10/28/94

It is virtually impossible, and highly impractical, to build an effective
magnetic shielding system for an entire laboratory room. About the best that
you can do is to locate the room in which the electron microscope will be
housed as far as possible from potential sources of large alternating
magnetic fields, such as: power lines that carry large currents;
transformer substations; building switch boxes and junction boxes; elevator
motors; etc. It is also very important to be sure that all metal water and
sewer pipes, and metal heating and ventilating ducts in the neighborhood of
the instrudment room are individually grounded some distance before they
enter the region of the lab, and that the electrical contractors are alerted
to the problems that can arise from large 'ground loop' currents flowing
through such items. Ground loops through steel structural members of the
building itself can also produce large magnetic fields. You need to talk to
the university architects and plant design engineers to be sure that every
possible precaution is taken to avoid them, too. You will have to be very
persistent about these matters, because most of the time construction people
are not concerned with them. It is also a very good idea to have a separate
ground rod installed in each instrument laboratory, so that each instrument
has a good electrical ground that is free from outside effects. Good luck!





From: kleifer-at-cimesg1.epfl.ch (Klaus Leifer)
Date: Fri, 28 Oct 1994 19:46:50 +0000
Subject: SAD precision 2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bob,

The strain of the substrate (we measure macrosopical stresses in the order
of 1e9 dynes/cm2), when calibrating the distances on the diffraction
pattern with the Si substrate reflections , surely is a problem in the case
of my specimen. Do you perhaps have citations concerning the relation
between substrate-film strain and corresponding lattice parameter
variations of the substrate?

Your idea to measure the distortions from high resolution images can be an
interesting alternative to the diffraction analysis.

Concerning the variations of 0.5% of spot positions in the diffraction
image: I first acquired a diffraction pattern, then changed to image mode
to look at the specimen. Afterwards I changed back to diffraction mode
again and made a second photo of a Si diffraction pattern. The evaluation
of both patterns showed deviations of the Si reflection positions between
both patterns of about 0.5% (comparing of course the same Si reflections,
not having touched diffraction focus and having defocussed the beam in both
cases the condensor until the end). I measured differences of 0.2-0.6%
between the reflection positions of the two negatives. All deviations went
in the same direction and the Si spots are sharp.

Concerning the calculation:
I wanted to know an order of magnitude of the deviation of the reflection
position, when one starts to tilt cristallites. In the the first
calculation I made two weeks ago I used the following basic geometry:

I
I e-beam directio=
n
I
I

______________________________________
I
I I foil thickness d
______________________________________
So the grain here has a thickness d in electron beam direction and is
infinitisemally long. I calculated then with the program of Pierre
Stadelmann the position of the diffraction spot position with respect to
tilt angles of the grain ( you also could calculate this analytically) and
the distance between the knot in the reciprocal space and the Ewald
sphere. From this distance (supposing a two beam case) I calculate the
intensity of the diffraction spot for every tilt angle of the grain.
This gives me a curve 'position of reflection/intensity' which is roughly a
(sinx)2/x2. So this curves would represent the form of a peak of a Dedye
Scherrer diagram with random orientation of grain with the shown geometry.
=46rom the width of this peak I can see, at which tilt angle reflections of =
a
grain are still visible and the corresponding shift of the reflection
position. I made this calculation for 1,2,5 and 10 nm thick grains and in
this geometry the possible deviation of the reflection positions from the
untilted grain was for 1nm thick grains was around than 1.5e-3, for 5nm
thick grains about 4e-4. Grains that are tilted by higher angles, this
means which show higher deviations in the reflection position are for a
factor 100 lower than the reflection at 0=B0tilt.
Of course, there would still be missing a treatment of grains with a
realistic form! Did somebody calculate this?
Best wishes
Klaus


__________________________________________________________________
Klaus Leifer, Ecole Polytechnique Federale de Lausanne (EPFL)
Address: EPFL-CIME, CH-1015 Lausanne, Switzerland
Phone: +41(21)693 48 30 Fax: +41(21)693 44 01
__________________________________________________________________






From: ldm-at-apollo.numis.nwu.edu (L. D. Marks)
Date: Fri, 28 Oct 1994 15:56:33 -0500
Subject: SAD accuracy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I noticed that while discussing SAD accuracy, noone mentioned
the fact that dynamical diffraction can change the position of spots:
see Hirsrch et al 1965; Hashimoto et al Phil Trans Roy Soc London
253 (1961) 459, or for a very short summary Marks, Ultramicroscopy 12
1984 237.




From: Wen-Shan Liou :      liou-at-acsu.buffalo.edu
Date: Fri, 28 Oct 1994 19:20:14 -0400 (EDT)
Subject: "FOCUS ON MICROSCOPY 95" call for papers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



FOCUS ON MICROSCOPY 95

A joint meeting of
8th International Conference on 3D Image Processing in Microscopy and 7th
International Conference on Confocal Microscopy

April 18-20, 1995
Howard Plaza Hotel, Taipei, Taiwan, Rep. of China



CALL FOR PAPERS


_________________________________________________________________

SCIENTIFIC PROGRAM

Scientific sessions start on Tuesday, April 18, 1995. Original
contributions will be presented in the following areas:

Confocal microscopy

* theory of confocal imaging
* scanning confocal designs: point / slit arrangements, beam
scanning, direct field scanning (bilateral, tandem)
* high resolution optical 3-D Microscopy
* two photon and time resolved fluorescence imaging
* transfer functions and deconvolution

Applications

* confocal microscopy in-vivo: approaches, problems and prospects
* 3D imaging for agricultural research
* fluorescence 3-D imaging in cell biology and neurobiology
* fluorescent probes, in-situ hybridization
* 3-D cytometry
* microstructures of materals, polymers and thin films
* application in environmental sciences

Near-field microscopy

* near-field scanning techniques (NSOM, STM, AFM)
* high resolution DNA - imaging
* spectroscopy and surface modification
* combined near-field and confocal designs

Optical tweezers and scalpel

* instrumentation
* applications

Electron Microscopy

* cryo-microscopy
* low-voltage SEM
* electron beam tomography

X-ray microscopy

* theory and instrumentation of x-ray microscopy
* x-ray sources

3-D imaging processing

* 3-D reconstruction of histological, optical and tomographical
sections
* visualization models in 3-D and 4-D microscopy
* supercomputing in microscopy
* analysis of serial section images, 3-D scene recognition
* 3-D image restoration and image quality

CONTRIBUTION AND PUBLICATION OF EXTENDED ABSTRACTS

Papers are invited for oral and poster presentation from the fields
indicated by the scientific program and related areas. Deadline for
submission of abstracts is December 31, 1994. An extented abstract
(minimum 700 words and maximum 1500 words) is required for each
presentation, and will be published as a suplement issue of Zoological
Studies (ISSN 1021-5506). All text will be typeset by the publisher.
Authors are encouraged to submit their manuscript in electronic forms.
Files created from the following word processos are acceptable,
otherwise, please submit your manuscript in ASCII form (both Mac and
IBM-PC format). Manuscript can also be submitted by e-mail to: elepcc-at-
ubvms.cc buffalo.edu, however, a hardcopy has to be sent by mail (or
faxed) to the address in USA. A hardcopy is required accompany the
electronic form.

Mac Word, Wordstar, Word Perfect, Microsoft Word, Ventura, ASCII

Figures and photos are permitted, however they have to fit the
following format specified in the photo and diagram format guide for
direct photoreproduction. Original photographs (both B&W and color)
and line drawings are required. If it is possible, authors please
provide a FAX number to facilatate the transmission of galley proof in
early 1995. The organizing committee will make a selection of the
abstracts for oral presentation. By the end of January 1995 authors
will be notified about acceptance and the final program will be mailed
to all registrants.

ACCOMMODATION

The Howard Plaza Hotel provides subtantially reduced room rates for
conference participants of Focus on Microscopy '95. Hotel
Accommodation at the Howard Plaza Hotel is offered on a first come,
first served basis. Please refer to Focus on Microscopy '95 for
qualifying the reduced room rates at booking: (refer to Registration
form)

Howard Plaza Hotel
160 Jen Ai Road, Sec. 3
Taipei, Taiwan, Republic of China

Phone: 886-2-700-2323
FAX: 886-2-700-0729

The room rate includes 10% service charge, welcome wine, fruit basket,
newspaper and the use of health club facilities including sauna. Major
credit cards (American Express, Visa, Mastercard, JCB, Diners Club)
are accepted at the hotel.

REGISTRATION & CONFERENCE FEE

The conference fee is US$220, which includes, documentation, abstract
book and refreshments during breaks. A preregistration fee of
US$180.00, is available when postmarked before January 31, 1995.

OFFICIAL AIR CARRIER

China Airlines is the official air carrier of Focus on Microscopy 95,
special discount airfare is available through CAL's world wide branch
offices.

INFORMATION

Registration, abstract forms and enquires:

N. America and Europe:

Focus on Microscopy '95 c/o Dr. P. C. Cheng
Advanced Micrscopy and Imaging Laboratory
Department of Electrical and Computer Engineering
State University of New York at Buffalo
P.O. Box. 84
Getzville, NY 14068
USA
Tel and Fax: 716-645-3868
e-mail: elepcc-at-corn.eng.buffalo.edu

Other nations:

Focus on Microscopy '95 c/o Dr. J. L. Wu
Institute of Zoology
Academia Sinica
Nankang, Taipei, Taiwan 11529
Republic of China
Tel: 886-2-789-9500
Fax: 886-2-789-9503/886-2-785-8059
e-mail: zojlwu-at-ccvax.sinica.edu.tw

ORGANIZERS


C.P. Chen (Taipei) G.J. Brakenhoff (Amsterdam) A.Kriete (Giessen)
C.H. Chou (Taipei) P. C. Cheng (Buffalo)(Chairman) C.J.R. Sheppard (Sydney)

P.P. Hwang (Taipei) C. Cogswell (Sydney) D.M.Shinozaki (London,Cana
da)
W.Y. Lee (Taipei) M. Gu (Sydney) E.H.K. Stelzer (Heidelberg
)
H.K. Wu (Taipei) V. Howard (Liverpool) T. Wilson (Oxford)
J.L. Wu (Taipei) H. Kim (Rochester)
W.L. Wu (Taipei)




_________________________________________________________________

THE CONFERENCE IS JOINTLY O RGANIZED AND SUPPORTED BY


The Society for 3-D Imaging Sciences in Microscopy, Amsterdam
Institute of Zoology, Academia Sinica, Taiwan, R.O.C.
Electron Microscopy Society of China, Taipei, R.O.C.
Life Science Research Promotion Center, NSC, R.O.C.
AMIL, State University of New York at Buffalo, U.S.A.


_________________________________________________________________


Wen-Shan Liou

ECE Dept. SUNY at Buffalo




From: imgp-at-mbimp1.mbl.TNO.NL (Kees van der Wulp)
Date: Sat, 29 Oct 1994 13:49:51 +0100 (MET)
Subject: Re Mag-shield for labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Glenn,

I completely agree with the message of Wil Bigelow that shielding a whole
building is not the right way if it would work at all.
If you will shield your measuring and recording equipment you can place
them together with the object you measure in a so called cage of Faraday.
There should be guidelines on how to build such a cage and of what material,
but it might be fairly expensive. In that case you might have a good shielding.

If you want to make precautions during construction time, not only look for
good grounding points seperated from the mains but also try to avoid
high temporal (far less than a second e.g. starting mercury arc lamps etc.)
current 'pulses' enter your room by using a saturated kind of transformer.
This kind of transformer (1:1) will block electrical pulses that may disturb
your measurements. I hope I made clear what I meant because I don't know
the english technical term.

If you use a number of wall plugs take care that this supply points all
have the same electrical phase (as you may know 220 volts is one of the 3 phases),
because different phase mains plugs can cause unwanted electrical signals
if you connect your instuments to them.

Hope you can use this information.

Regards
--
Kees van der Wulp
TNO - Nutrition & Food Research Institute INTERNET : vanderwulp-at-mbl.tno.nl
Division of Toxicology VOICE : +31 15 843101
PO-Box 5815 FAX : +31 15 843989
2280 HV RIJSWIJK (NL)
THE NETHERLANDS





From: ard-at-atom.ansto.gov.au (Arthur Day)
Date: Sun, 30 Oct 1994 01:17:04 +1000
Subject: Re Mag-shield for labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {aad8145d010210048093-at-[137.157.15.210]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Glenn,

A few years ago a company had set up a stand trying to sell magnetic
shielding at one of our local microscopy conferences. We were setting up
new labs, so I asked them about it. They were selling "boxes" generally
used for keeping high intensity fields in, rather than low intensity fields
out, for enclosing NMR machines etc. The dimensions in their illustrations
were not much larger than the size of an SEM, although I suppose they could
build bigger boxes. The walls were pretty thin-looking and when asked they
said it was a "special material" :-). Somehow I don't think that putting
one of these around a microscope would do much good, since low intensity
50-60 Hz magnetic fields are not really easy things to block.

Their price for a "box"? A mere $45,000 AUD. :-)

Arthur Day, Electron Microscopy Group
Ansto Advanced Materials Program Phone: 61-2-717-3457
PMB 1, Menai (Sydney), NSW, 2234 Fax: 61-2-543-7179
Australia
Email: ard-at-atom.ansto.gov.au







From: A. Kent Christensen :      akc-at-umich.edu
Date: Sat, 29 Oct 1994 17:18:04 -0400 (EDT)
Subject: Re: Re Mag-shield for labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I can't add any suggestions to what others have said about magnetic
shielding for labs. However, the need for shielding reminds me of several
years (1971-78) I spent at Temple University School of Medicine, located
on Broad Street in north Philadelphia, Pennsylvania. One of the city's
main subways went underground up Broad Street. I remember someone's EM
facility at ground level in a small building. Every time a subway would
come by, the EM beam would undergo rather striking excursions on the
screen. They tried various approaches (Faraday cage, Gaussian ring,
etc.), but I don't think it was ever fully solved.

Our EM facility on the 6th floor of the medical school building
seemed all right. However, an EM newly installed in another department on
our floor had terrible intermittent beam shifting problems. It was hard
to see how it could be due to the subways, since our EMs seemed OK. After
a few bewildering weeks (both for the company and for the investigator) it
was discovered that the EM had been installed near an old (but still
operational) elevator that had an extremely heavy iron counterweight, that
traveled up and down all day long. Whenever the counterweight passed the
6th floor, the EM beam drifted off to the side, then drifted back. Who
would have thought of anything like that?




From: ldm-at-apollo.numis.nwu.edu (L. D. Marks)
Date: Sun, 30 Oct 1994 11:52:36 -0600
Subject: Re Mag-shield for labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


As a small comment, it can be just as effective to shield only
critical parts of the column using Nu-metal. Using one graduate student,
a pair of metal cutters and a local company for Hydrogen annealing of the
pieces you can perform miracles for $200-$300.




From: jjmill-at-RMIT.EDU.AU
Date: Mon, 31 Oct 1994 08:03:37 EST-10
Subject: re:mag shielding for labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


It may be of interest to know that at least one company (Link, through
Oxford) offers a magnetic shielding system which is an active
feedback system to compensate for generated fields in the laboratory.
I am not sure whether it is just for 50/60 ac or dc or both, and I have
no experience of the system, but would be mildly interested to hear
anyone's experience.
I agree with the comments regarding prior checks and appropriate
location rather than shielding.
Cheers
jjm
Professor John J. Millar
Head, Department of Applied Physics
Electron Microscope Unit
RMIT, GPO Box 2476V, MELBOURNE AUSTRALIA 3001
+613 660 2602 fax 613 660 3837
email jjmill-at-rmit.edu.au




From: attmail!internet!AAEM.AMC.ANL.GOV!MICROARCHIVE
Date: Mon Oct 31 10:00:31 -0600 1994
Subject: another

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Original-Subject: another
Original-Date: 0 Oct 31 10:00 -0600 1994
Confirming-MTS-Message-ID: {internet3041602030}
Not-Delivered-To: rratleng!internet!rratleng1!rratlengp0!allsman-at-attmail.com due to 01 Invalid Address Specification

Content-Type: Text

The AT&T Gateway has experienced error(s) delivering your message to
the Microsoft Mail System. The message may or may not have been
routed to the recipient. All errors encountered are listed below.

An error was found with the primary Recipient.
Address -} rratleng!internet!rratleng1!rratlengp0!allsman



Content-Type: Text
Content-Length: 2037

From uucp Mon Oct 31 16:02:01 GMT 1994 remote from attmail
} From AAEM.AMC.ANL.GOV!MICROARCHIVE Mon Oct 31 16:00:19 +0000 1994 remote from internet
Message-Version: 2
} To: rratleng!internet!rratleng1!rratlengp0!allsman
by with SMTP ID 0r1z9f.136BrU;
Mon, 31 Oct 1994 16:00:25 +0000
MTS-Message-ID: {internet3041602030}
UA-Content-ID: {941031110127.712-at-AAEM.AMC.ANL.GOV}
End-of-Header:
EMail-Version: 2




From: attmail!internet!AAEM.AMC.ANL.GOV!MICROARCHIVE
Date: Mon Oct 31 10:00:50 -0600 1994
Subject: another

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Original-Subject: another
Original-Date: 0 Oct 31 10:00 -0600 1994
Confirming-MTS-Message-ID: {internet3041602180}
Not-Delivered-To: rratleng!internet!rratleng1!rratlengp0!allsman-at-attmail.com due to 01 Invalid Address Specification

Content-Type: Text

The AT&T Gateway has experienced error(s) delivering your message to
the Microsoft Mail System. The message may or may not have been
routed to the recipient. All errors encountered are listed below.

An error was found with the primary Recipient.
Address -} rratleng!internet!rratleng1!rratlengp0!allsman



Content-Type: Text
Content-Length: 1669

From uucp Mon Oct 31 16:02:12 GMT 1994 remote from attmail
} From AAEM.AMC.ANL.GOV!MICROARCHIVE Mon Oct 31 16:00:46 +0000 1994 remote from internet
Message-Version: 2
} To: rratleng!internet!rratleng1!rratlengp0!allsman
by with SMTP ID 0r1zA6.13PRHk;
Mon, 31 Oct 1994 16:00:47 +0000
MTS-Message-ID: {internet3041602180}
UA-Content-ID: {941031110205.712-at-AAEM.AMC.ANL.GOV}
End-of-Header:
EMail-Version: 2




From: smithj-at-acad.winthrop.edu
Date: Thu, 25 Aug 1994 20:00:54 -0400
Subject: Coates & Welter *stubs* --free to good home

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


As part of a surplus purchase of SEM specimen boxes, I have
a few dozen used aluminum stubs that appear to go with Coates &
Welter. Having an ISI/Topcon 'scope, I have no use for them.
If you'd like them mailed to you, please contact me directly.
Please *do not* reply to the list.
Julian Smith III
Dept of Biology
Winthrop University
Rock Hill, SC
803-323-2111 (vox)
803-323-2246 (fax)
smithj-at-winthrop.edu





From: attmail!internet!AAEM.AMC.ANL.GOV!MICROARCHIVE
Date: Mon Oct 31 09:59:18 -0600 1994
Subject: Here's one of them

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Original-Subject: Here's one of them
Original-Date: 0 Oct 31 09:59 -0600 1994
Confirming-MTS-Message-ID: {internet3041559470}
Not-Delivered-To: rratleng!internet!rratleng1!rratlengp0!allsman-at-attmail.com due to 01 Invalid Address Specification

Content-Type: Text

The AT&T Gateway has experienced error(s) delivering your message to
the Microsoft Mail System. The message may or may not have been
routed to the recipient. All errors encountered are listed below.

An error was found with the primary Recipient.
Address -} rratleng!internet!rratleng1!rratlengp0!allsman



Content-Type: Text
Content-Length: 1461

From uucp Mon Oct 31 15:59:42 GMT 1994 remote from attmail
} From AAEM.AMC.ANL.GOV!MICROARCHIVE Mon Oct 31 15:59:17 +0000 1994 remote from internet
Message-Version: 2
} To: rratleng!internet!rratleng1!rratlengp0!allsman
by with SMTP ID 0r1z8f.11yLRI;
Mon, 31 Oct 1994 15:59:17 +0000
MTS-Message-ID: {internet3041559470}
UA-Content-ID: {941031110037.712-at-AAEM.AMC.ANL.GOV}
End-of-Header:
EMail-Version: 2




From: Walt Bobrowski :      bobroww-at-aa.wl.com
Date: Mon, 31 Oct 1994 13:02:10 -0400 (EDT)
Subject: Lead Citrate pH

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Is anyone routinely checking pH of prepared lead citrate according to Bozzola
(12.0 +/- 0.1)?






From: masroor-at-engrs.unl.edu (Masroor Malik)
Date: Mon, 31 Oct 1994 12:57:34 -0600
Subject: Lead Citrate pH

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everybody,
I am a regular reader of all the mails that are in Microscopy.
I really admired people willing to help each other.
Now I have a problem I need some input from all of you experienced
guys out there. I am working on mechanically alloyed materials and
it has become really hard to make samples out of it on quick basis
It usually takes a long time to embed it in epoxy and mechanically
polish and then Ion mill.

Now we have decided to explore a different path and are trying
to press the powder in a die till about 30000lbf. We have started to
work on Copper so we can optimize the technique and use it for other
materials as well. I am experiencing difficulties keeping the
samples powder in tact and it keep breaking on me when I try to cut
the sample out of pressed material.
If somebody has worked on it or has some knowledge about the
process I would really appreciate any help.

Masroor
masroor-at-engrs.unl.edu






From: COOK-at-AAEM.AMC.ANL.GOV
Date: Mon, 31 Oct 1994 16:03:53 -0600 (CST)
Subject: EM shielding/ M. Rich

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since your building plans are not final, now is the time to make the
appropriate measurements for magnetic fields. To do this job thoroughly, you
should lay out the proposed areas in a grid and then measure the horizontal
magnetic fields in 3-dimensions. This procedure will enable you to site your
electron columns properly or, if the fields are too large everywhere within the
grid, take preventative action before the building plans are approved and the
contracts are signed.




From: Liang, Long :      LLIANG-at-is.Arco.COM
Date: 31 Oct 1994 14:53:14 CST
Subject: AC magnetic fields

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MACMS.LLIANG.090459140094304FMACMS-at-IS.ARCO.COM}

Could someone teach me how to convert the unit mV (millivolt) to mG
(milligauss)? Recently we asked one service engineer to survey our new
EPMA/SEM lab. He wrote down the readings of AC magnetic fields in mV.

Thanks in advance.

Long Liang
Electron Microprobe and SEM Lab
ARCO Exploration and Production Technology
2300 West Plano Parkway
Plano, TX 75075
E-mail : LLIANG-at-is.Arco.COM







From: IAN HALLETT :      ihallett-at-marccri.marc.cri.nz
Date: Tue, 1 Nov 1994 10:29:50 GMT+1200
Subject: Re: EMSCOPE address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Date sent: Fri, 28 Oct 94 10:05:55 EDT
} From: dfalcon-at-VNET.IBM.COM
} To: microscopy-at-aaem.amc.anl.gov
} Subject: EMSCOPE address

} I have an EMSCOPE model SC650 large sample sputter coater that needs
} some repair parts. The phone number for the manufacturer in England that is
} on the paperwork that came with the unit is no longer a valid number. Does
} anyone know if the company (EMSCOPE LABORATORIES LTD--KENT,ENGLAND) is still
} in existence and what their present number is---and/or any North American
} representatives?? I suspect they may have been bought by another company
} and may have a new name and number now.
} Thanks,
} Doug


Doug

Your suspicion is correct and EMSCOPE no longer exists, they were
bought out some years ago by Biorad (Polaron) who later sold most of
their microscopy business in the UK (and elswhere) to VG Technologies
and Fisons. They may still suport some of the old EMSCOPE equipment.
Another alternative is to contact a firm called EMITECH which was
founded by some ex members of the EMSCOPE staff and they may be able
to help you with your problems.

US address: EMITECH INC, 3845 FM, 1960-West, Suite 345, Houston,
Texas 77068. Fax 713-893-8443. Phone 713-893-2067 or 800-444-3137
toll free for sales and service.

UK address: 11 Enterprise Centre, Newtown Road, Ashford, Kent, TN24
0PD. Fax 0233 640744 Phone 0233 646332

I have no connection to any of the above firms.

Ian








Ian









Ian Hallett
HortResearch
Mt Albert Research Centre
Private Bag 92 169
Auckland, New Zealand
Fax 64-9-815 4201
Telephone 64-9-849 3660




From: gkennedy-at-ucsd.edu
Date: Mon, 31 Oct 1994 13:37:41 -0800
Subject: lead citrate pH

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199410312134.NAA25918-at-ucsd.edu}
X-Sender: gkennedy-at-popmail.ucsd.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I've checked mine with a narrow-range paper--seems to be pretty close, about 11.8 or so. I tried adjusting it, but saw no difference in staining. Grace Kennedy, UCSD






From: M.Dickson-at-unsw.edu.au (Mel Dickson)
Date: Mon, 31 Oct 1994 11:14:09 +0000
Subject: Re: Microscope labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


--- Forwarded message follows ---

Hello Glenn,
We had a fearful problem with 150Hz (3 phase) VLFI in our laboratory. Partly
due to the fact that the laboratory was located adjacent to a major power
transformer substation for the university. I spent some years working on the
problem trying various solutions. First I got a search coil and plotted the
fields. There is a commercial one made here called a "Gauss Maus" You can get
them cheaply now there is paranoia about health effects of VLFI. A big
surprise was that that the radiation was maximal up against the wall of the
substation but it did not fall away as the square of the distance inside our
buiding. Instead it was high near the walls, vertical columns and cross
bearers in our steel framed concrete building. We did an experiment turning
power off to all sections of the faculty a bit at a time and found that field
was proportional to total power use and not to any one section of the building
which would have implicated particular cabling runs. My conclusion was that
the field was being induced in the frame of the substation which was welded
into the frame of the main building so they were magnetically coupled. Then
the frame of the whole building and every bit of reinforcing joined to the
frame had the field induced in it. I found almost the same field throughout
the building no matter how far from the substation. Of course you can also
get local fields generated in heavy cables but
A. live and return cables are usually routed side by side and the fields
cancel each other
B. the field falls off with square of distance so unless the cable goes right
over your column the field is negligible.
You can also get fields induced in airconditioning ducts water pipes etc. and
you might need to insert non metallic sections to isolate sections in the e.m.
lab.
A local firm RFI Industries 54 Holloway Drive Bayswater Melbourne VIC 3153
Australia built a very effective fully screened room for the E.M. in a nearby
hospital. They got good attenuation using quite cheap sheet metal and not the
really pricey magnetic shields like hipernoom. By the way many engineers
imagine it only takes chicken wire to shield a room. This is OK for radio
frequencies but entirely useless for 50/60 Hz. We tried local shielding of the
column but it made no difference. All columns have shielding built in already.
And unless you make a full metal box which completely surrounds the column
you really cant keep VLFI out. It just sneaks in the holes you have to leave
to work the controls!
We tried using a cancelling field which worked fine in
trials but couldn't cope in the microscope itself. AC VLFI fields are complex
multi directional 3D interference patterns and to compensate them fully you
need compensating fields on 3 axes. Also the sensor detecting the field to
generate the reverse phase feedback needs to be as close as possible to the
volume you wish to be compensated. Since the microscope itself generates
fields you will pick up e.g. the SEM scan signals and fed them back so as to
neutralise them.
There is a book in the Series " Practical Methods in Electron Microscopy" Ed.
Audrey Glauert. Volume 4, "Design of the Electron Microscope Laboratory" by
R.H. Alderson. North-Holland Publishing Co., American Elsevier, N.Y. ISBN
0-444-10807 6 PUBLISHED 1975. Which deals with this and other problems in a
very practical way.
In the end we solved our problem by moving to a low field building. Hope you
manage better!
Mel Dickson.






From: Daniel E. Sampson :      des-at-rupture.ucsc.edu
Date: Mon, 31 Oct 94 16:27:48 CST