From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Tue, 1 Nov 1994 08:42:46 -0500
Subject: frog skin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Having processed some frog skin for TEM, looking at the thick sections I
noticed the tissue looks like small balls. Can anyone tell me what this
might be? Looking at normal skin such as human tissue this looks completely
different. Is there a good book that might describe what these structures
are?
TIA,
Phil
8-{)




From: OPUS::EMLAB 1-NOV-1994 08:46
Date: Tue, 01 Nov 1994 08:52:24 -0400 (EDT)
Subject: TEM perfusion of rat tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



----- Begin message from OPUS::EMLAB 1-Nov-94


Karen,

We routinly use 4% paraformaldehyde, 2% glutaraldehyde in 0.2M sodium
cacodylate buffer at pH 7.2. We probably need a mininium of 500 ml of fixative
to perfuse the rat. After perfusion, store whole organs in 3% glut or
dice up tissue and store in buffer. Good Luck.

Ed Calomeni


----- End forwarded message






From: nee-at-lanl.gov (Norman Elliott)
Date: Tue, 1 Nov 1994 07:19:43 -0700
Subject: gas regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear listreaders,

We are in the process of having installed a new JEOL TEM and the
service engineer suggested we use an "on-demand" gas regulator to control
the dry nitrogen vent gas for the column. This type of regulator allows
the pressure in the column to reach 1 atmosphere but no more to prevent
over pressuring the column and blowing out vacuum seals, thin windows, etc.
I know I have seen such regulators but I cannot find a source. If anyone
is using this type of regulator could you please supply a model number and
vendor.


Thanks in advance.




Norman Elliott | E-mail: nee-at-lanl.gov
Los Alamos National Lab | Fax: 505-665-2104
MST-7 MS E549 | Voice: 505-667-1587
Los Alamos, NM 87545 |






From: Jay Jerome :      jjerome%isnet-at-ncren.net
Date: Tue, 1 Nov 1994 10:11:27 -0500 (EST)
Subject: Re: gas regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We could not find an inexpensive regulator from gas supply houses, etc.
so we use regulators designed for scuba tanks. Still not cheap, but less
expensive than the alternaives we found. They have worked well without
problem for about 8 years now.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************



On Tue, 1 Nov 1994, Norman Elliott
wrote:

} Dear listreaders,
}
} We are in the process of having installed a new JEOL TEM and the
} service engineer suggested we use an "on-demand" gas regulator to control
} the dry nitrogen vent gas for the column. This type of regulator allows
} the pressure in the column to reach 1 atmosphere but no more to prevent
} over pressuring the column and blowing out vacuum seals, thin windows, etc.
} I know I have seen such regulators but I cannot find a source. If anyone
} is using this type of regulator could you please supply a model number and
} vendor.
}
}
} Thanks in advance.
}
}
}
}
} Norman Elliott | E-mail: nee-at-lanl.gov
} Los Alamos National Lab | Fax: 505-665-2104
} MST-7 MS E549 | Voice: 505-667-1587
} Los Alamos, NM 87545 |
}
}
}







From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Tue, 1 Nov 1994 09:16:02 -0600
Subject: Wehnelt apertures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I recently replaced the Wehnelt aperture in my Philips CM12 electron gun. The
original aperture had a hole diameter of 0.6 mm and I replaced it with one that
has a 0.8 mm dia. A 0.4 mm hole is also available in my column kit.

I know from my experience on an SEM that increasing the hole diameter will
increase the beam current output from the gun (which I did to get more counts
for x-ray microanalysis).

What else is affected by the hole size - resolution, coherence, or what? How
should one decide what diameter to use for this aperture on an SEM or a TEM?

--

Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu






From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 1 Nov 1994 10:13:02 U
Subject: Re: EM shielding/ M. Rich

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1428437492.33680-at-mse.engin.umich.edu}

Reply... RE} EM shielding/ M. Rich
Reply
from: John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX: (313)936-3352
jfmjfm-at-umich.edu or John_Mansfield-at-mse.engin.umich.edu
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html

Opinions and experiences here are my own.

We have a field problem in our JEOL 4000EX room. Before all the microscopes
were installed the JEOL engineers did a field survey in each room. All was
fine. They then installed a 2000FX in one room and got it up and running and
then in the adjacent room installed the 4000. Now both reached spec without
problem. But, as many of you may know, JEOL had some fearful problems with
the guns of the 4000 series. We had really bad filament image flickering and
it never seemed to be totally cured (in spite of extensive glow discharging,
oops! I mean conditioning).
Tracing the problem and trying to blame everything except the gun, lead our
engineers (and there were several groups of them!) to test for fields and
they notice that there was a substantial field about 5 feet from the scope in
the floor. It was below the required level at the objective lens, but that
didn't stop them from mentioning it (many times). Turns out that the supply
for the 2000 ran under the floor of the 4000 room and the phases were
sufficiently unbalanced to produce this field.
Now, the 4000 scope had been installed after the 2000 was up and running and
so we had taken the resolution tests with that field present and so we didn't
think it was a problem. This did not prevent the engineers trying to say
that our gun problem was related to that field. The alternative at this
point was to admit that the gun was bad and replace it. After a good deal
of discussion the gun was replaced. By this time the scope had been down so
long that I had asked JEOL Service for a rebate on my service contract since
they had not maintained the scope in operating condition for a reasonable
portion of the period covered by the contract.
The moral of this story is. When you build your scope facility, make sure
that you specify where everything goes. And I mean everything, water, power,
ground lines, light locations (fluorescent lights can interfere with the
scopes, the chokes have quite a field)., everything. If we hadn't had the
field in the floor we would not have had to argue about it! It's easier if
you don't build in problems for yourself!
Don't let the builders or the bean counters change the plans after you have
approved them. Our plans were amended and we now have building air instead
of our own system and I always have toasting microscopes in the fall and
spring when the A/C is in the process of being turned on or off!
As I say just my opinion and experiences. I am not trying to get at JEOL, it
is just that it took a long time to convince them that it was their gun and
not our facility.

John Mansfield.





From: Damon Heer :      DLH-at-fei2.feico.com
Date: Tue, 1 Nov 1994 07:55:34 -800
Subject: LaB6 Fix

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings to all on the listserver.

The following is an abbreviated version of an offline email dialogue
between myself and Paul Nolan of Queen's University regarding his
"LaB6 Fix" question. Paul, Nestor and I agreed that this would be a good
dialogue to share with the listserver. If anyone has any questions
regarding these issues, feel free to contact me.

Best regards,
Damon L. Heer.
***************************************************************

I'VE GOT A LaB6 THAT STOPPED EMITING. LOOKS GOOD UNDER THE SEM
EXCEPT FOR THE TUNGSTEN BALLS ALL OVER IT! GOING TO TRY TO CLEAN IT
IN MY CARBON EVAPORATOR. THINK IT WILL WORK? ANY OTHER IDEAS? (CAN'T
AFFORD A NEW ONE)

THANKS, PAUL
-----------------------------------------------------
Dear Paul:

You noted that it looks good in the SEM. Do you mean that the
crystal still appears to have a reasonable emitting surface
remaining? how long did it run before it stopped emitting?

You also noted a build up of tungsten balls on it. Is this build up
on the mount or the crystal? I've never seen this kind of
contamination on a tip. this is because I mainly deal with our LaB6
cathodes which don't have tungsten in the structure. However, if you
are using a Denka cathode, the cement that is used to keep the
crystal in the mount may be causing this contamination.

To clean it, we sometimes rub the crystal with a q-tip with a
little isopropyl alcohol. If this doesn't remove it, try lightly
rubbing it with the wooden stick portion of the q-tip, again with
IPA. This basically applies a rather non-abrasive scrub to the
crystal. The difficulty here is that both Denka's and Kimball's LaB6
mounts are very fragile and may break under any contact.

Unfortunately, Ed Calomeni is very correct in saying that your best
bet is buying a new one.

Good luck!
Damon L. Heer
------------------------------------------------------
Damon.
There seemed to be no structural damage to the filament ie cracks or
defects but the tungsten balls cover the whole filament.

The filament has been in our TEM for about 2 years but has not had
too many hours of use . It is a Kimball filament so i dont know if
there is any W in the mount or not. I will give it a wipe and look
at it in the sem again and see what has happened. In your opinion is
this W going to stop the crystal from emitting? (There appears to
be no other damage and there is continuity between the mounting
posts) ...

Cheers, Paul
-------------------------------------
Hi Paul,

In my opinion, if there is no contamination on the cone and flat of
the crystal, it should work fine. When an electron microscope is
saturated properly, the emission only comes from the flat area.

You noted that the cathode stopped emitting while in use. The reason
this usually happens is the emitting flat at the tip of the crystal
cone has evaporated to a point. If there still is a flat left on
the emitter, I imagine that the flat did become contaminated, but is
unnoticeable in the SEM.

I know that a large amount of time is necessary for changing tips in
a TEM, so the following recommendation should be weighed against that.

If there remains a reasonable amount of the flat on the crystal, and
the W can be cleaned away from the crystal cone and flat, it may be
possible to drive off any remaining contamination by operating the
cathode at about 0.2 amps above the normal operating filament current
for a short period of time. This may recover the cathode.

I hope things work out.

Cheers,
Damon
---------------------------------------
Damon. Thanks again.

You mentioned that the flat on the filament was the important part to
have clean, well it seems there is no definite flat part on the tip.
I was unaware because of my inexeperience with LaB6 that there should
be a flat, i assumed it should be pointed (like mine seem to be) Is
this filament dead?

cheers paul
-----------------------------------
Hi Paul,

Kimball, as well as FEI, uses a "truncated tip" for LaB6 emission.
The truncated tip consists of a cone with a flat top that is normal
to the optical axis. The "flat" is oriented with the {100} crystal
plane, which has been shown to be the best orientation for electron
microscope applications. The cone angle and the flat diameter are
chosen by the type of application it will be used for. FEI's
standard cone angles and flat diameters (and I believe the same goes
for Kimball) are as follows:

SEM and Low Res TEM
90 degree full cone angle, 16 micron flat diameter

Hi Res TEM
60 degree full cone angle, 5 micron flat

When the flat is gone, the emission comes from the side of the cone,
which isn't the {100} plane and is not normal to the optical axis. The
result is poor emission down the column.

If there is no flat left on your tip, I'd say it's dead.

Coincidentally, a colleague and I just wrote an article for our
semi-annual newsletter, the FEI Focus, regarding crystal tip issues.
If you wish, send me your address and I could send you a copy of
this. A reprint also appears in the most recent issue of Microscopy
Today.

Cheers,
Damon

FEI Company
7451 N.E. Evergreen Parkway
Hillsboro, OR 97124-5830

Phone (503) 640-7582
fax (503) 640-7509
email dlh-at-feico.com




From: W.L. Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Tue, 1 Nov 1994 13:14:28 EST
Subject: Gas Regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Norman,

Last year, we installed a JEOL JEM 1210 and were advised to purchase
the demand flow regulator for the dry nitrogen. We ended up with a
Matheson Model 3421 vacuum regulator which meets JEOL's specs...in fact
its one that they recommend. Its pricey at about $400, and is available
from any of the industrial gas suppliers, ie AIRCO, etc. However, its
maximum inlet pressure is 50 psi, meaning that you can't couple it
directly to 3000 psi nitrogen tank. You still must buy a general purpose,
dual stage nitrogen regulator, then attach the demand flow regulator to
this. The whole setup will end up costing about $600.
Incidentally, if you are installing a TEM, then you know that you will
have to buy a cylinder of sulfur hexaflouride and a regulator for this. A
100 lb bottle of sulfur hex is about $400, but you can use a nitrogen
regulator for its installation if you buy a $15 adaptor. I can supply
this information if you need it. Good Luck.

-=buddy=-

-=W.L. Steffens=-
College of Veterinary Medicine
University of Georgia




From: randy nessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Tue, 1 Nov 1994 13:14:02 -0600 (CST)
Subject: SEM of chocolate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have someone interested in the fracture surfaces of choclate. I
want to use some type of replication process, as I don't want to put
chocolate in my SEM. No, I don't want ESEM, or cryoSEM. I am worried
that the solvents of formvar or similar plastics might distort the
fracture surface. I need low mag shots, so high resolution replicas are
not necessary.
thanks in advance,
Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu









From: randy nessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Tue, 1 Nov 1994 12:59:48 -0600 (CST)
Subject: Re: Gas regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I deleted the original post on this subject, but from what I
remember, I think the original poster was looking for a pressure relief
valve that tripped at 1 atm, instead of a regulator that would supply 1
atm. We used to have a little pop-off valve that came with a thin window
x-ray detector. It was supposed to be installed in the line that bleeds
the chamber, so as to prevent a positive pressure from forming in the SEM
specimen chamber. A positive pressure ran the risk of rupturing the
window. You might call Kevex/Fisons, since this is who supplied the
pop-off valve for us.
Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu








From: Greg Erdos ICBR EM Core Lab Univers :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Tue, 01 Nov 1994 14:52:00 -0500 (EST)
Subject: SiO and AlO films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can anyone point me in the right direction for techniques for making silicon
monoxide and aluminum oxide films. A discussion of the resultant surface prop-
erties would also be helpful, if such info is written down someplace.

Thanks
**********************************************************
* Greg Erdos ** *
* Director, ICBR EMCL ** Phone 904-392-1295 *
* 218 Carr Hall ** FAX 904-846-0251 *
* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
* Gainesville, FL 32611 ** *
****************************************************************




From: tivol-at-tethys.ph.albany.edu
Date: Tue, 01 Nov 1994 16:30:50 EST
Subject: Wehnelt apertures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Gib,
The wehnelt aperture size and the fields produced in and around the
filament and wehnelt determine the locus of points from which electrons emitted
by the filament end up being accelerated and sent down the column. The larger
the aperture size, 1) the more electrons go down the column, 2) the less cohe-
rent the beam, 3) the larger the crossover size, and 4) the more the anode is
heated by electrons extracted at angles significantly different from 0 degrees.
The best wehnelt aperture size depends on what the instrument is used
for. The largest possible size is that for which anode heating becomes a pro-
blem. For people like me who do low-dose work and diffraction and whose EDX
detector gives high dead time for more intense beams, a smaller aperture is
better; for those who are fighting to increase intensities all the time, larger
is better.
Yours,
Bill Tivol




From: EMLAB-at-opus.mco.edu
Date: Tue, 01 Nov 1994 16:48:14 -0400 (EDT)
Subject: Re: gas regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We use such a device, which was purschased a local scuba diving supply
store. They work great.

Ed Calomeni




From: M.Dickson-at-unsw.edu.au (Mel Dickson)
Date: Wed, 2 Nov 1994 07:08:56
Subject: Re: gas regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-aaem.amc.anl.gov

In article nee-at-lanl.gov (Norman Elliott) writes:
}
} Dear listreaders,

} We are in the process of having installed a new JEOL TEM and the
} service engineer suggested we use an "on-demand" gas regulator to control
} the dry nitrogen vent gas for the column. This type of regulator allows
} the pressure in the column to reach 1 atmosphere but no more to prevent
} over pressuring the column and blowing out vacuum seals, thin windows, etc.
} I know I have seen such regulators but I cannot find a source.


} Hi Norman. The regulator you need is available at a cost of about $200 from
dive shops selling scuba gear. They are just breathing demand valves which
allow gas to flow when you suck on the valve but not otherwise. Its important
to always have gas available on the pressure side or the microscope vacuum
will suck the valve diaphragm to destruction! We have such a system feeding
our SEM from the bled off gas in out 175Litre Liquid nitrogen tank, which is
the driest gas you can get. If you have a UTW window on your EDS you might
install a relief valve on the line so pressure in the microscope won't ever go
over atmospheric.

Good luck, Mel.




From: W.L. Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Tue, 1 Nov 1994 18:18:28 EST
Subject: Sulfur hexafluoride

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sulfur hexafluoride is the insulating gas used in high voltage tanks. In
the early days, all tanks were filled with dielectric oils, and some of
them contained PCB, forcing the manufacturers to find another type oil.
As accelerating voltages increased, the manufacturers switched to Freon as
a superior insulatant. Of course recently, the EPA has banned the use of
most Freons, so now most manufacturers have switched to sulfur hex., which
is supposed to be completely inert. The HV tanks don't come pressurized
with it, so you have to supply it for the service engineer. As I
understand it, SEMs with their lower accelerating voltages still have oil
in the tanks.

-=W.L. Steffens=-
College of Veterinary Medicine
University of Georgia




From: Fred Pearson :      eoptics-at-mcmail.cis.mcmaster.ca
Date: Tue, 1 Nov 1994 16:14:50 +0001 (EST)
Subject: Re: Wehnelt apertures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Tue, 1 Nov 1994, Gib Ahlstrand wrote:

} I recently replaced the Wehnelt aperture in my Philips CM12 electron gun. The
} original aperture had a hole diameter of 0.6 mm and I replaced it with one that
} has a 0.8 mm dia. A 0.4 mm hole is also available in my column kit.
}
} I know from my experience on an SEM that increasing the hole diameter will
} increase the beam current output from the gun (which I did to get more counts
} for x-ray microanalysis).
}
} What else is affected by the hole size - resolution, coherence, or what? How
} should one decide what diameter to use for this aperture on an SEM or a TEM?
}
} --
}
} Gib Ahlstrand, MMS Newsletter Editor
} Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
} 495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
} 612-625-9728 FAX, giba-at-puccini.crl.umn.edu
}
}
We also have a CM12 with LaB6 filament. The general "rule of thumb" is to
multiply the Wehnelt distance, ie. distance you screw back the filament
into the anode, by 2 1/2 to obtain the Wehnelt aperature size you require.
I back off the filament by 200 um. Therefore I require a 500 um. aperature.
The lifetime of the filament is just under 1000 hrs.

Hope this gives you some assistance.

Fred Pearson
Electron Optics Facility
Institute for Materials Research
McMaster University
Hamilton, Ontario
voice (905) 525-9140 x24609
fax (905) 521-2773
eoptics-at-mcmail.cis.mcmaster.ca





From: Jan Coetzee EM Univ Pretoria :      JANC-at-ccnet.up.ac.za
Date: Wed, 2 Nov 1994 08:18:33 GMT+2
Subject: Re: SEM of chocolate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

microscopy-at-aaem.amc.anl.gov

} I have someone interested in the fracture surfaces of choclate. I
} want to use some type of replication process, as I don't want to put
} chocolate in my SEM. No, I don't want ESEM, or cryoSEM. I am worried
} that the solvents of formvar or similar plastics might distort the
} fracture surface. I need low mag shots, so high resolution replicas are
} not necessary.
} thanks in advance,
} Randy Nessler
} rnessler-at-emiris.iaf.uiowa.edu

What will most probably work in your case is one of the dental replicating
silicones. These are formulated to make quick, accurate replicas of tooth
surfaces and are used in the preparation of ceramic (and other)
crownwork. Any dentist should be able to help. We have used these silicones
in replicating leaf surfaces, as well as human skin, and the resolution is OK
up to at least 1000x mag. There are other (cheaper) types of dental
impression materials available, but these water-based formulations give low
resolution and are of course not suitable for direct exposure to the SEM
vacuum. Have fun.

Jan Coetzee
Unit for Electron Microscopy Tel:+27-12-420-2075
University of Pretoria Fax:+27-12-342-1738
Pretoria Internet:janc-at-ccnet.up.ac.za
0002 South Africa




From: RETEP-at-anat.uct.ac.za
Date: 2 Nov 94 08:36:19 SAST-2
Subject: Re: Frog Skin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {MAILQUEUE-101.941102083619.1088-at-anat.uct.ac.za}
To: microscopy-at-aaem.amc.anl.gov

In answer to Phil's question regarding frog skin:

What area of the skin are you looking at? If it is only the
epidermis, the 'balls' are probably the epidermal cells which have
shrunk (a common occurence) during fixation. If it is dermal then
you may well be looking at glands situated in the dermal region. If
the balls are within the cells then the possibility is that they are
the frog equivelant of melanin.

I am unaware of any decent book on frog skin per se. I do have some
references which you may find useful if you wish to have them.

Peter

_______________________________________________________________


-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
-at- -at-
-at- Peter D. G. Richards -at-
-at- Dept Anatomy and Cell Biology -at-
-at- UCT Medical School -at-
-at- Observatory -at-
-at- 7925 -at-
-at- RSA -at-
-at- Tel: 021-406 6285. -at-
-at- Internet: retep-at-anat.uct.ac.za -at-
-at- -at-
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-




From: Richard E. Edelmann :      REDELMAN-at-musom01.mu.wvnet.edu
Date: Wed, 02 Nov 1994 08:44:03 +1100
Subject: Sulfur hexaflouride?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



O.k., I'll ask for us ignorant folks. Why is a tank of sulfur
hexaflouride needed for the installation of a new TEM as was
mentioned in responding to the low cost regulator series?


Richard E. Edelmann
Electron Microscopy Facility Supervisor
Marshall University - School of Medicine
Huntington, West Virginia




From: imgp-at-mbimp1.mbl.TNO.NL (Kees van der Wulp)
Date: Wed, 02 Nov 1994 13:22:52 +0100 (MET)
Subject: Astromed Liquid Nitrogen CCD-3200 systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello fellow microscopists out there,

I am interested to get into contact with those who use an
Astromed liquid nitrogen CCD camera, type 3200, on top of
any type of microscope to quantify fluorescence images.
I would like to exchange experience and test images.

Please mail me directly at : vanderwulp-at-mbl.tno.nl

Hope to hear from you.

PS. Does anybody know of a mailing list used by astronomers
because this equipment is frequently used by them.

Thanks for any comment.
--
Kees van der Wulp
TNO - Nutrition & Food Research Institute INTERNET : vanderwulp-at-mbl.tno.nl
Division of Toxicology VOICE : +31 15 843101
PO-Box 5815 FAX : +31 15 843989
2280 HV RIJSWIJK (NL)
THE NETHERLANDS





From: Richard E. Edelmann :      REDELMAN-at-musom01.mu.wvnet.edu
Date: Wed, 02 Nov 1994 08:44:03 +1100
Subject: Sulfur hexaflouride?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1428347675.69421-at-mse.engin.umich.edu}

Reply... RE} Sulfur hexaflouride?
Reply
from: John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX: (313)936-3352
jfmjfm-at-umich.edu or John_Mansfield-at-mse.engin.umich.edu
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html

On a realted note, how are people managing the switch from Freon to SF6?
According to our JEOL Service engineers, the upgrade of our 4000 to SF6 is a
freebie as part of our maintenance contract, but they want to stiff us for
$20K to upgrade the 2000FX. I have been hoarding a few large cylinders of
freon to make sure we can continue to run and service our 2000 for a number
of years with freon. What are other folks doing?
Jfm.

--------------------------------------

O.k., I'll ask for us ignorant folks. Why is a tank of sulfur
hexaflouride needed for the installation of a new TEM as was
mentioned in responding to the low cost regulator series?


Richard E. Edelmann
Electron Microscopy Facility Supervisor
Marshall University - School of Medicine
Huntington, West Virginia

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;2 Nov 1994 10:00:12 -0400
Received: by truelies.rs.itd.umich.edu (8.6.9/2.2)
with X.500 id IAA00456; Wed, 2 Nov 1994 08:58:32 -0500
Received: from AAEM.AMC.ANL.GOV by truelies.rs.itd.umich.edu (8.6.9/2.2)
with SMTP id IAA00450; Wed, 2 Nov 1994 08:58:30 -0500







From: ANLAEM::CSENCSITS 2-NOV-1994 12:36:11.94
Date: Wed, 2 Nov 1994 12:36:56 -0600 (CST)
Subject: EM Shielding and the JEOL 4000

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Our JEOL 4000 is affected by magnetic fields due to any TV monitor in
the room that is not very new and highly shielded. This includes Mac
monitors and the TV monitor for our TV rate camera.

The filament appears to flicker at 3-5 Hz and we are unable to measure
a magnetic field at that frequency even near the monitor let alone the
microscope! We do measure the 50-60 Hz field at the monitor and the
stronger this field the larger the effect on the beam. The effect is
greater the closer the monitor is to the scope and the closer to the
objective lens.

For best images we turn off the monitors before exposing film. There is
an obvious difference!! For some seeing was believing.

Roseann Csencsits
Argonne National Laboratory
Argonne, Illinois 60439
USA
708-252-4977




From: Jay Jerome :      jjerome-at-isnet
Date: Tue, 1 Nov 1994 10:11:27 -0500 (EST)
Subject: Re: gas regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



this is a reposting- our mail program has not been sending out all
correspondence and that which it sends out has had the wrong return
address. Sorry for cluttering cyberspace if you've seen this before.

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************

---------- Forwarded message ----------

We could not find an inexpensive regulator from gas supply houses, etc.
so we use regulators designed for scuba tanks. Still not cheap, but less
expensive than the alternaives we found. They have worked well without
problem for about 8 years now.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************



On Tue, 1 Nov 1994, Norman Elliott
wrote:

} Dear listreaders,
}
} We are in the process of having installed a new JEOL TEM and the
} service engineer suggested we use an "on-demand" gas regulator to control
} the dry nitrogen vent gas for the column. This type of regulator allows
} the pressure in the column to reach 1 atmosphere but no more to prevent
} over pressuring the column and blowing out vacuum seals, thin windows, etc.
} I know I have seen such regulators but I cannot find a source. If anyone
} is using this type of regulator could you please supply a model number and
} vendor.
}
}
} Thanks in advance.
}
}
}
}
} Norman Elliott | E-mail: nee-at-lanl.gov
} Los Alamos National Lab | Fax: 505-665-2104
} MST-7 MS E549 | Voice: 505-667-1587
} Los Alamos, NM 87545 |
}
}
}








From: Fred Pearson :      eoptics-at-mcmail.cis.mcmaster.ca
Date: Thu, 3 Nov 1994 10:47:28 +0001 (EST)
Subject: OPTICAL DIFFRACTOMETERS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Our lab is considering building, or purchasing (new or used) an Optical
Diffractometer.
Does anyone have a list of suppliers, a used instrument, or wishes to
share their design?
Comments on manufactured models are welcomed.
Email directly if you wish to avoid biased opinions.

Thanks
Fred Pearson
fax (905) 521-2773
McMaster University
Hamilton Ontario, Canada
eoptics-at-mcmail.cis.mcmaster.ca





From: pbiesiot-at-whale.st.usm.edu (Patricia M Biesiot)
Date: Thu, 3 Nov 1994 11:46:54 -0600 (CST)
Subject: LM: video/microscopy info needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I need advice, recommendations, information about video microscopy
systems; I want to buy a system to use in my undergraduate classes (gen.
zool., invert. zool.) to observe small, live animals. I especially need
specific brands, models that others have used successfully in similar
situations. The students themselves will be using the system and making
and editing videos--so the equipment needs to be fairly sturdy yet enable
good quality images. At this time I have price quotes from a number of
sources (Carolina Biological, Fisher, Nikon, Olympus) but of course each
salesperson recommends his/her system! Thanks for any advice.
P.M. Biesiot
Dept. Biol. Sci.
Univ. Southern Mississippi
Hattiesburg, MS 39406-5018
pbiesiot-at-whale.st.usm.edu




From: Garth Freeman :      garth.freeman-at-gtri.gatech.edu
Date: Thu, 3 Nov 1994 15:50:15 -0600 (CST)
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English

unsubscribe:garth.freeman-at-gtri.gatech.edu
I am leaving Georgia Tech to join Materials Analytical Services in
Norcross Georgia.




From: Calvin Montgomery :      cal-at-marlin.ssnet.com
Date: Thu, 3 Nov 1994 16:22:31 -0500 (EST)
Subject: Re: Sulfur hexaflouride?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Wed, 2 Nov 1994, Richard E. Edelmann wrote:

}
} O.k., I'll ask for us ignorant folks. Why is a tank of sulfur
} hexaflouride needed for the installation of a new TEM as was
} mentioned in responding to the low cost regulator series?
}
SF6 is used in only some new TEMs. Hitachi uses SF6 in the guns of all
TEMs that go to 200KV and over. And as of yet only in the HV tanks of the
FE-TEMs. Its used as an insulator, instead of oil or Freon.
Cal-Hitachi service




From: M.Dickson-at-unsw.edu.au (Mel Dickson)
Date: Fri, 4 Nov 1994 09:17:02
Subject: Re: OPTICAL DIFFRACTOMETERS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-aaem.amc.anl.gov

For years we used a basic optical bench with a low power laser and lenses we
scrounged from microscopes and enlargers and cameras. The best idea is
to buy two 2 meter benches and use an optically flat mirror (astronomy
supply house stock) to bend the pattern back so you can view the pattern as
you move the negative around. The best viewing/recording device is an old
4x5 camera of some sort. You don't even need one with a working shutter or
light tight bellows as you will be using the diffractometer in the
dark. You can then easily use polaroid 45 materials for hard copy of the
patterns. BUT you will have to be cautious about viewing the pattern and
always e.g. have the zero order beam blocked with a metal patch. I have the
scars to show how careful you need to be.

email me for further information. Mel Dickson
} Thanks } Fred Pearson} fax
(905) 521-2773} McMaster University} Hamilton Ontario,
Canada} eoptics-at-mcmail.cis.mcmaster.ca







From: Daniel Luchtel :      dluchtel-at-u.washington.edu
Date: Thu, 3 Nov 1994 16:40:45 -0800 (PST)
Subject: Re: frog skin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: dluchtel-at-homer18.u.washington.edu

The "small balls" structures might be mucous glands. See fig. 5 in Hauser
F et al 1990. Expression of spasmolysin (FIM-A.1): an integumentary mucin
from Xenopus laevis. Exp Cell Res 189: 157-162 - it shows a phase
constrast micrograph of Xenopus skin with mucous glands that might be
interpreted as "small balls".

On Tue, 1 Nov 1994, rutledge phil wrote:

} Having processed some frog skin for TEM, looking at the thick sections I
} noticed the tissue looks like small balls. Can anyone tell me what this
} might be? Looking at normal skin such as human tissue this looks completely
} different. Is there a good book that might describe what these structures
} are?
} TIA,
} Phil
} 8-{)
}




From: DRK%SHCC.BITNET-at-ANLVM.CTD.ANL.GOV
Date: Thu, 3 Nov 1994 17:44 PSTT
Subject: Subscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe DRK-at-shcc.bitnet





From: OLINSAL-at-BIOAX1.BIO.ORNL.GOV
Date: Fri, 4 Nov 1994 7:19:12 -0500 (EST)
Subject: DISSECTING MICROSCOPES

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,
I have been approached by a first grade teacher at Pond Cove
Elementary School who wants to instill an early love for microscopy in
the children. Are there any free or low-cost dissecting microscopes
out there??
Thanks, Ada
Ada Olins
Univ. Tennessee-Oak Ridge GSBMS
Biology Division, ORNL
P.O. Box 2009
Oak Ridge, TN 37831-8077

Tel: 615 574 1269
Fax: 615 574 1274
e-mail: olinsal-at-bioax1.bio.ornl.gov




From: Greg Erdos ICBR EM Core Lab Univers :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Fri, 04 Nov 1994 08:55:59 -0500 (EST)
Subject: Methylamine Tungstate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Michel et al.,
I have used methylamine tungstate quite successfully for seing some
very fine filaments on the surface of tooth decay bacteria. I follow Wyatt
et al., 1988. Oral Microbiol Immunol 3:162-168.
If this journal is not widely available outside dental schools just
send a psotal address or fax # and I can get a copy to you.

Greg
******************************************************
* Greg Erdos ** *
* Director, ICBR EMCL ** Phone 904-392-1295 *
* 218 Carr Hall ** FAX 904-846-0251 *
* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
* Gainesville, FL 32611 ** *
******************************************************************




From: A. Kent Christensen :      akc-at-umich.edu
Date: Fri, 4 Nov 1994 09:14:30 -0500 (EST)
Subject: Re: Sulfur Hexafluoride

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Being a scientist doesn't insure good spelling, any more than being
in a company insures courtesy.

A. K. Christensen

---------------------------------

On Thu, 3 Nov 1994, John Mardinly, 5-2346, Page 322-6490, SC2-24 wrote:

} Folks;
} What I want to know is why there are so many of you out there claiming
} to be scientists yet can't get the spelling of fluoride any better than
} "flouride"?
}
} John Mardinly
} Intel MATTEC
}




From: tvoiles-at-unlinfo.unl.edu (todd voiles)
Date: Fri, 4 Nov 1994 08:31:26 -0600 (CST)
Subject: site for SuperSem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A few weeks ago a person posted a note about having a
SuperCard stack about SEM's and offered the FTP site for
it........I replyed to him and asked for the site name
but never got any reply.....could somebody help me out?
It would be great for teaching and training.....

Thanks

--

E.Todd Voiles *** Facility Manager
Central Facility for Electron Microscopy
University of Nebraska
tvoiles-at-unlinfo.unl.edu




From: Marcelle A Gillott :      magem-at-csd.uwm.edu
Date: Fri, 4 Nov 1994 08:47:24 -0600 (CST)
Subject: gas regulators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hve you tried a pressure relief valve?

Our Hitachi has a "branch" in the vent line with a pressure relief valve
which then allows us to use a standard regulator - would that be a
suitable solution to the problem?






From: COOK-at-aaem.amc.anl.gov (Russell E. Cook)
Date: Fri, 4 Nov 1994 10:03:57 -0600
Subject: Optical diffractometer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We have used a Polaron optical diffractometer for several years. It
came as a complete package: bench, lenses, laser, holder for
negatives, etc. The addresses were:

Polaron Equipment Ltd.
53-63 Greenhill Crescent
Watford Business Park
Watford Hertfordshire WD1 8QS

Polaron Instruments, Inc.
2293 Amber Drive
Line Lexington Industrial Park
Hatfield, PA 19440
(215)822-2665

Russell E. Cook
Electron Microscopy Center for Materials Research
Materials Science Division
Argonne National Laboratory
Argonne, Illinois 60439
USA






From: Marcelle A Gillott :      magem-at-csd.uwm.edu
Date: Fri, 4 Nov 1994 09:04:10 -0600 (CST)
Subject: Re: TEM - Negative Staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hve you tried bring the pH of the uranyl up to around neutral? this is
what i generally do when using it as a "negative" stanin

another alternate to try is PTA (phosotungstic acid)

good luck
marcelle gillott





From: MicroToday-at-aol.com
Date: Fri, 4 Nov 1994 10:08:59 -0500
Subject: Students/Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good Friday Morning,
Ada Olins (Univ. of TN) asks for (inexpensive) microscope sources suitable
for young students.
In this broad area, I would like to introduce the "DiscoveryScope" to
readers. It is "an inexpensive, exceptionally durable viewing system for
exploring small things. It's 1-inch focal length lens produces beautiful
wide field, high resolution views of the small invertebratres, tiny plants,
protist, and inorganic subjects. Special chamber and holders are provided
for viewing both terrestrail and aquatic organisms. Not just for beginners,
DiscoveryScope makes a useful "spotting scope" for any microscopist who
collects samples from the field, particularly if macro photography is
intended."
Discovery Scope Inc
PO Box 607
Green Valley, AZ 85622
Tel/Fax: (800)398-5404
I have no financial interest in this product (unfortunately) - and but
submit that it is a nifty device!
Don Grimes, Microscopy Today





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Fri, 4 Nov 1994 11:05:47 -0600 (CST)
Subject: SuperSEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



If all goes well SuperSem will be posted on the

Microscopy & Microanalysis Public Domain Software Library site
in a few weeks. Brendon Griffin will be visiting ANL in November
and will be bringing a copy along. The FTP site is:

WWW.AMC.ANL.GOV

It currently hosts the EMMPDL, MASLIB, Test Images, and PDL Software
all for Microscopy&Microanalysis.


Your Friendly Neighborhood SysOp (who can sometimes spell)

Nestor




From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 4 Nov 1994 15:26:22 -0400
Subject: SuperSEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English

John Mansfield
North Campus Electron Microbeam Analysis Laboratory
413 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX (313)936-3352
jfmjfm-at-engin.umich.edu or John.F.Mansfield-at-umich.edu Time:
14:24

Date:11/4/94
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html NC EMAL

I currently have a working copy that Brendon sent me and I will make it
available when he gives me the OK. Warning, it is an 11 megabyte file and
requires at least 5 megabytes to run. It runs only on a Mac.





From: COOK-at-aaem.amc.anl.gov (Russell E. Cook)
Date: Fri, 4 Nov 1994 15:38:17 -0600
Subject: re nitride etching

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We were able to etch Si3N4 (silicon nitride) in a CF4 plasma using a Plasma
Prep II from SPI Supplies. See: Mat. Sci. & Eng., A169 (1993) L5-L7.

Russell E. Cook
Electron Microscopy Center for Materials Research
Materials Science Division
Argonne National Laboratory
Argonne, Illinois 60439
USA






From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Fri, 4 Nov 1994 16:33:06 -0500 (EST)
Subject: lookin for a cheap shutter for Nikon Diaphot light source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anybody have a suggestion for a shutter for a Nikon Diaphot light source?
We're looking for the following requirements:
- easy to mount on upper (halogen) or lower (mercury) lamp input
- computer controlled via NIH-Image (or IP-Lab Spectrum)
- cheap!

The two main applications are:
- time lapse w/ fluorescence
- collecting random fields w/ fluorescence

At this point we don't need selectable filters, just a shutter, although
we wouldn't mind getting various filter positions as part of the deal.

Thanks-
Michael Cammer cammer-at-aecom.yu.edu






From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Fri, 4 Nov 1994 16:50:28 -0500 (EST)
Subject: TIFF images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


anybody know an FTP site with an MSDOS program for displaying TIFF
files? Archie and veronica turn up lots of pictures but no programs.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sat, 5 Nov 1994 13:06:21 -0600 (CST)
Subject: PC Viewers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jay Jerome asked about TIF viewers for PC machines...

There are currently NO TIFF viewers in the ANL Library for PC's.
but the ANL FTP site has GIF viewers for PC machines (both DoS and WinDoze)
the are in the following path:

ANLAEM Software Library
ANL Shareware Library
Freeware Shareware PC
Viewers

The site address is: WWW.AMC.ANL.GOV
The programs are called: VGIF.exe, VUGIF.exe, WINGIF.exe, CSHOW.exe

The PC commerical program PhotoStyler certainly opens TIFF files, and
so do a lot of other commerical programs (PhotoShop etc...).

As a note of warning. Some PC TIFF reading programs will not read TIFF
files that were created on a MacIntosh. Most Mac programs that I've
used, on the other hand, are intelligent enough to recognize the differences
and translate appropriately. Also not all TIFF readers can handle
the full TIFF specification. Many are coded assuming that 8 bit
gray scale and/or 24bit (8x3) color is the end product.
If you have 12,16 ... bit scientific data many (but not all) will
not display properly. Make sure you take this into account if your
looking at something other than photographic quality images.

There are also the usual suite of Translator programs which will convert
between the different image formats. Those programs are usually cheaper
than the full blown image processing/editing programs. By purchasing
one of these translator programs you could then use the Shareware Viewers....

Your Friendly Neighorhood SysOp...

Nestor






From: Tina Carvalho :      tina-at-ahi.pbrc.Hawaii.Edu
Date: Sat, 5 Nov 1994 17:37:06 -1000 (HST)
Subject: Hardware needed to run NIH IMAGE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We have been using a 286 PC-based Kontron IBAS-II with IBAS 2.0 software
for image analysis for several years with only middling success.

We am now assisting some researchers with semi-quantitative analysis of
the silver grains from their in situ hybridization studies. They are
interested in purchasing a Macintosh system to run the NIHIMAGE program
you have developed. Their interest primarily stems from the following
paper: Lucas, L.R., Mize, R.R. and Harlan, R.E. Semiquantitative
analysis of in-situ hybridization results using IMAGE software: a rapid
method for counting reduced silver grains over mRNA-positive cells. J.
Neuroscience Methods 52:101-109 (1994). They have asked us to help them
determine what computer and video hardware they need to set up their
system.

Is anyone willing to suggest the required hardware for an adequate Mac
system to run NIHIMAGE? We want neither a Volkswagen nor a Cadillac, just
a good level system with a Macintosh they can also use for other
purposes. Do you recommend a Quadra or a PowerPC? Which model? We
assume that 8 MB RAM is minimum and 16 MB is better. Do you have any
particular video camera and frame grabber your recommend? (We currently
have a DAGE MTI CCD72 video camera). Our computer network manager has a
nubus VideoSpigot frame grabber available if it would be of use. We
assume it is best to run the program using two monitors. Any
recommendations for the video monitor?

We've noticed from reading some of the information on the microscopy usenet
newsgroup that a PowerPC version of NIHIMAGE is being developed. Is the
PowerPC version tested enough for a "novice" user? We have about 5 years
of image analysis experience on the IBAS system so we know much of the
terminology and the steps involved in image analysis. We just son't have
a clue about the Mac...

Any assistance you can provide would be greatly appreciated.

Aloha,
Tina Weatherby Carvalho
Marilyn F. Dunlap
Biological EM Facility
Pacific Biomedical Research Center
University of Hawaii at Manoa
1993 East-West Road
Honolulu, HI 96822
(808)956-6151 or (808)956-6251 or FAX (808)956-4768
e-mail: dunlap-at-ahi.pbrc.hawaii.edu
tina-at-ahi.pbrc.hawaii.edu







From: M.Dickson-at-unsw.edu.au (Mel Dickson)
Date: Mon, 7 Nov 1994 09:41:58
Subject: Re: cheap magnetic shielding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-aaem.amc.anl.gov

In article

}
} { {Michael Rich, rich-at-egr.msu.edu
} { {Michigan State University
asks about cheap magnetic shielding..
Michael, there is no such thing because of the physics of magnetic fields.
Radio frequency interference CAN be excluded by making a Faraday cage - a
complete box - using copper mesh or even fine chicken wire mesh. BUT to
excludelow frequency magnetic fields you must use solid sheets of some
ferromagnetic material. Any gaps between sheets must be covered over by
strips of similar material. To entirely shield a room the six faces of the
room must have the shield applied (yes that includes the floor and ceiling).
The material must either be a high permeability material like the alloy
hipernom or it may be ordinary steel sheet but much thicker. To shield a
complete room costs $40 - 50,000. Much of the cost lies in the fact that the
construction should be done just right to obtain best shielding. Try looking
up the yellow pages for a shielding supplier. I have an old address for one:
Amuneal Manufacturing Corp, 4737 Darrah St. Philadelphia Pa 19124.

All the best, Mel Dickson.

}




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Mon, 7 Nov 1994 17:18:30 +1300
Subject: PVP solubility

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A question from Allan Mitchell

Recently I have been using PVP (polyvinylpyrrolidone) in various techniques
but to date I have had no luck.
Firstly I have tried PVP with LR Gold resin (as recommended by the LR
technical blurb), I have also tried adding it my fixative to increase
osmotic pressure (as recommended in the paper I have been following to
localise the enzyme Thiamine Pyrophosphatase) and finally I have tried it
when infusing tissue with cryoprotectant for cryoultramicrotomy (as
recommended by Tokuyasu)

In these papers they talk of "dissolving" the PVP in the various solutions
but my PVP will not dissolve!!! I am left with a fine white suspension
which quickly settles out when stirring is stopped.

Question: Should the PVP dissolve (and perhaps leave a clear solution) or
is the fine suspension I have normal? (It blocks up canulas when perfusing,
it makes it difficult to see the tissue when processing).

Note: The Merck index states that PVP is soluble in alcohols and water -
have I got the wrong one?

Many thanks in anticipation.

Allan Mitchell

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: Petr Schauer :      petr-at-ISIBrno.Cz
Date: Mon, 7 Nov 1994 12:06:10 +0100
Subject: Re: TIFF images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} anybody know an FTP site with an MSDOS program for displaying TIFF
} files? Archie and veronica turn up lots of pictures but no programs.

Jay,
Try Graphic Workshop (gws70c.zip), which is located for example at the
ftp.uni-mannheim.de (134.155.50.51): /disk2/systems/msdos/graphics

Regards,

Petr
+-------------------------------------------+-------------------------+
| Dr. Petr Schauer | tel.: (+42 5) 41321246 |
| Head of Electron Microscopy Laboratory | fax : (+42 5) 41211168 |
| ACADEMY OF SCIENCES OF THE CZECH REPUBLIC | E-mail: petr-at-isibrno.cz |
| INSTITUTE OF SCIENTIFIC INSTRUMENTS | petr-at-mrkev.vabo.cz |
| Kralovopolska 147, CZ-612 64 Brno | |
| Czech Republic | |
+-------------------------------------------+-------------------------+




From: _ _ _ _ _ MARK W. LUND _ _ _ _ _ :      LUNDM-at-PHYSC1.BYU.EDU
Date: Mon, 7 Nov 1994 12:12 MST
Subject: Sulfur hex

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {Chameleon.941107102153.tonygr-at-emlab.mit.edu}

We used tons of sulfur hexafluoride when I supervised a lab testing image
intensifiers. We would flood bare tubes with it, so breathed a lot of it.
It is non-toxic, and can only make you pass out if it excludes the oxygen.
Sulfur hex actually quenches sparks as it decomposes in the arc. As for
fluorine remaining in the tank this is highly unlikely, since it would react
with anything it came into contact with.

As for spelling, I have come to the rule of thumb for fluorine, fluorescence
etc. that if it spells flour it is wrong. I still have fond feelings for
a professor who gently pointed this out to me.

regards
Mark W. Lund
MOXTEK, Inc.
Orem UT




From: Alan Pooley :      pooley-at-ahab.rutgers.edu
Date: Mon, 7 Nov 1994 14:48:49 -0500
Subject: Re: frog skin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The balls seen might be 'membrane blisters' as written up in scanning vol 1
p 166-173 1978 E. Shelton & Mowczko
the basic idea is that OsO4 sometimes creates '? blebs 'like little balloons
on the surface of cells, these were up to 6 microns long but mainly 1-2 microns
or sometimes smaller *(they say .2 to .6) in a variety of cells and
conclude that glutaraldehyde a lone is best foir fixation.
Since many people have not found this, it may only be certain tissues that do
this in most people's routine handling of fixation.
Alan Pooley Marine science rutgers nj




From: Alan Pooley :      pooley-at-ahab.rutgers.edu
Date: Mon, 7 Nov 1994 15:01:35 -0500
Subject: Re: Sulfur Hexafluoride

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Concerning spelling errors etc (flouride vs fluoride etc) may I suggest to
all the gentlemen and lady scientists that we quietly suggest corrections of
errors ti individuals and avoid blaring them over the net.
( I expect an individual correction for my mistyping of 'to' in the previous l
Alan Pooley marine sci sem lab rutgers nj




From: GREG VAUGHN MARTIN :      gmartin-at-welchlink.welch.jhu.edu
Date: Mon, 7 Nov 1994 15:00:54 -0500 (EST)
Subject: Re: PVP solubility

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hey folks: A reply to Allen Mitchell -- I think you're problem might lie
in the molecular weight of the PVP you're using, or the temperature
you're trying to dissolve it at, or both. We routinly use PVP with a
m.w. of 40,000 (Sigma cat. PVP-40) in fixatives for in situ perfusion of
rat livers -- 2% w/v takes a while (several minutes) to go into sol'n,
but it does go in. I also make a Tokuyasu-style infiltration medium for
cryosectioning -- that uses PVP m.w. 10,000 (Sigma PVP-10) but it is
dissolved into NaCO3-phospate (100g of PVP into 30ml of liquid!) and must
be done by alternating between a 60C oven and a sonicator full of hot
water. Feel free to E-mail me (gmartin-at-welchlink.welch.jhu.edu) for
particulars of preparing this stuff (it's a real witch's brew). I can't
speak as to the particulars of PVP added to LRGold -- I've always used
LRGold straight-up -- but I would advise against putting you're specimens
into a sol'n cloudy with undissolved PVP.

Take Care, Greg Martin
Dept Cell Biology and Anatomy
Johns Hopkins School of Medicine

On Mon, 7 Nov 1994, Richard Easingwood wrote:

} A question from Allan Mitchell
}
} Recently I have been using PVP (polyvinylpyrrolidone) in various techniques
} but to date I have had no luck.
} Firstly I have tried PVP with LR Gold resin (as recommended by the LR
} technical blurb), I have also tried adding it my fixative to increase
} osmotic pressure (as recommended in the paper I have been following to
} localise the enzyme Thiamine Pyrophosphatase) and finally I have tried it
} when infusing tissue with cryoprotectant for cryoultramicrotomy (as
} recommended by Tokuyasu)
}
} In these papers they talk of "dissolving" the PVP in the various solutions
} but my PVP will not dissolve!!! I am left with a fine white suspension
} which quickly settles out when stirring is stopped.
}
} Question: Should the PVP dissolve (and perhaps leave a clear solution) or
} is the fine suspension I have normal? (It blocks up canulas when perfusing,
} it makes it difficult to see the tissue when processing).
}
} Note: The Merck index states that PVP is soluble in alcohols and water -
} have I got the wrong one?
}
} Many thanks in anticipation.
}
} Allan Mitchell
}
} Richard Easingwood
} Department of Anatomy and Structural Biology,
} P.O. Box 913
} University of Otago,
} Dunedin, New Zealand.
} Fax:64-3-479 7254
} Telephone:64-3-479 7301
}
}
}




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Tue, 8 Nov 1994 09:56:26 +1300
Subject: Re: PVP solubility reference

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Reply to: RE} PVP solubility
}
} Richard,
} Could you send me the reference for the use of PVP, i.e. the Tokuyasu ref?
}
} Thanks---Mike
}
} Mike_Schwartz-at-qm.yale.edu
} Fax 203-785-5263
} Voice 203-785-4324
} Michael Schwartz, Ph.D.
} Associate Professor
} Section of Neurobiology
} Yale University School of Medicine
} 333 Cedar St.
} New Haven, CT 06510

Mike, the reference is Tokuyasu, K.T.(1989) Use of poly(vinylpyrrolidone)
and poly(vinyl alcohol) for cryoultramicrotomy.Histochemical Journal 21,
163-171. I will fax a copy to you if you like.

Richard Easingwood
Department of Anatomy and Structural Biology,
P.O. Box 913
University of Otago,
Dunedin, New Zealand.
Fax:64-3-479 7254
Telephone:64-3-479 7301






From: efosten-at-mmm.com
Date: Mon, 7 Nov 1994 17:51:50 -0600
Subject: Spelling errors / typos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


One last (?) word on spelling errors:
"It's a damn small mind that can think of only one way to spell a word."
- Andrew Jackson (seventh President of the U.S.)




From: Gerald Little :      ANGJL-at-medicine.newcastle.edu.au
Date: Tue, 8 Nov 1994 08:25:57 GMT+1100
Subject: PVP solubility - A Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} A question from Allan Mitchell
}
} Recently I have been using PVP (polyvinylpyrrolidone) in various techniques
} but to date I have had no luck.
} Firstly I have tried PVP with LR Gold resin (as recommended by the LR
} technical blurb), I have also tried adding it my fixative to increase
} osmotic pressure (as recommended in the paper I have been following to
} localise the enzyme Thiamine Pyrophosphatase) and finally I have tried it
} when infusing tissue with cryoprotectant for cryoultramicrotomy (as
} recommended by Tokuyasu)
}
} In these papers they talk of "dissolving" the PVP in the various solutions
} but my PVP will not dissolve!!! I am left with a fine white suspension
} which quickly settles out when stirring is stopped.
}
} Question: Should the PVP dissolve (and perhaps leave a clear solution) or
} is the fine suspension I have normal? (It blocks up canulas when perfusing,
} it makes it difficult to see the tissue when processing).
} Note: The Merck index states that PVP is soluble in alcohols and
water - } have I got the wrong one?
} Many thanks in anticipation.
}
} Allan Mitchell
}
} Richard Easingwood
} Department of Anatomy and Structural Biology,
} P.O. Box 913
} University of Otago,
} Dunedin, New Zealand.
} Fax:64-3-479 7254
} Telephone:64-3-479 7301

Dear Allan,
We make up sucrose/PVP as a cryo-protectant using the following
technique. It finallly goes into solution after being stirred
overnight.
Dissolve 30g PVP in 80ml water at approx 60C, add sucrose (79g) and
20 ml 0.5M PBS. Stir in coldroom overnight. Store at 4C. The
solution is a clear yellow. Much like a good Hunter Chardonnay.
Hope this helps.
Regards,
Gerald Little.

Dr Gerald J. Little | Ph (61 49) 215618
The Neuroscience Group |
Discipline of Anatomy | Fax (61 49) 216903
Faculty of Medicine and |
Health Sciences |
The University of Newcastle | Email ANGJL-at-medicine.newcastle.edu.au
Australia, 2308 |




From: David Brauer :      dbrauer-at-arserrc.gov
Date: Tue, 08 Nov 1994 05:49:55 -0400 (EDT)
Subject: calibration of snarf

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

MR-Received: by mta ISHTAR.MUAS; Relayed; Tue, 08 Nov 1994 05:49:55 -0400 (EDT)
MR-Received: by mta ISHTAR; Relayed; Tue, 08 Nov 1994 05:49:55 -0400 (EDT)
Disclose-recipients: prohibited

We have been using several different fluorescein based dyes to
measure pH gradient within epidermal cells of roots from maize
seedlings in situ. Treatment of the tissue with snarf-am under
the appropriate conditions leads to a rather uniform labeling
of both the vacuole and cytoplasm. Therefore, the snarf in
these cells should be in pH environments ranging from pH 5 in
the vacuole to almost pH 8 in to cytoplasm. We would like to
be able to measure this pH gradient. The pH dependence graph
of snarf in the Molecular Probes book makes one think it
should be able to detect pH changes throughout this range. We
have a calibration curve using a ratio of two excitation
wavelengths of 420 and 475 nm at a constant emission
wavelength of 580 + 30 nm. However, this ratio varies only
between pH 7 and 9. We have experimented with another
calibration curve using a ratio intensities with excitation
and emission wavelengths of 420 nm and 580 nm compared to
that obtained at exciation and emission wavelength of 480 and
620 nm. With standard solutions, this ratio shows pH-dependent
changes from pH 5 to 9 but the majority of the change is
between pH 7 and 8. However, the calibration does not look as
clean when one tries to do the calibration by clamping cells
at various pH by various means.
Any suggestions would be appreciated.
Thank You for Your Reply in advance,
Dave Brauer
plant Physioogist
dbrauer-at-arserrc.gov






From: RETEP-at-anat.uct.ac.za
Date: 8 Nov 94 14:19:14 SAST-2
Subject: Re: Recoating phosphor screens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411081155.AA22042-at-riker.ml.wpafb.af.mil}

Dear All,

Anyone out their re-coat their phosphor screens?

How does one go about it? Are the supplies available in South Africa?

The screen I am proposing to coat has an alluminium base, about 12 cm
diameter, for a Hitachi TEM.

Many thanks for any advice.

Peter Richards.

_______________________________________________________________


-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
-at- -at-
-at- Peter D. G. Richards -at-
-at- Dept Anatomy and Cell Biology -at-
-at- UCT Medical School -at-
-at- Observatory -at-
-at- 7925 -at-
-at- RSA -at-
-at- Tel: 021-406 6285. -at-
-at- Internet: retep-at-anat.uct.ac.za -at-
-at- -at-
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-




From: Richard E. Edelmann :      REDELMAN-at-musom01.mu.wvnet.edu
Date: Tue, 08 Nov 1994 08:36:33 +1100
Subject: Viewing Tiff images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


If your simply interested in viewing Tiff images (and many other
formats) on a PC I would suggest Norton Desktop for Windows (by
Symantec). The Norton Veiwer does offer limited image manipulation
such as zooming. If your like me and have basically been forced from
DOS to the Windows environment for $90 Norton Desktop greatly improves
windows and has alot more to offer than just viewing and shuffling
files (though as a replacment for Windows Filemanager alone I would
recomend it). I don't know if the lastest version of Norton
Commander has added Tiff viewing compatablity - but I wouldn't doubt
it - and it runs under DOS without windows.

If you need additional image manipulation or another "viewer"
WordPerfect 6.0 ($130 - educational discount) handles Tiff format
with no problem, and allows limited image manipulation (zooming, cut,
pasting, contrast,brightness) and printing. I would assume that most
other current word processors should have similar features.

But for "real" image processing you need to spend much more money
for the dedicated image processing programs. You might want to check
out the article "High-end Image Editing Software" in Computer
Shopper, October 1994, 14: 548-558, and the article "Graphical
Voodoo" in the same issue p.610-622 which offers an excellent review
of the various imaging formats pros and cons.


Richard E. Edelmann
Electron Microscopy Facility Supervisor
Marshall University - School of Medicine
Huntington, West Virginia




From: swatkins-at-pitt.edu (simon)
Date: Tue, 8 Nov 1994 08:50:07 -0500
Subject: Post doc opportunity

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Position available
A postdoctoral position is available within the Structural Biology Imaging
Center of the University of Pittsburgh Medical School. The focus of the
position will be to study the interactions/role of spectrin like proteins in
muscle, brain and other tissue systems. Specifically dystrophin and alpha
actinin. The principal approach will be to use a variety of high resolution
light and electron microscopic techniques (confocal, Immuno-EM, etc) to
define where, when, and in association with which proteins these specific
molecules and isoforms of these molecules are found. The applicant should
have experience/interest in modern microscopic methods and be willing to
learn other techniques as needed.

If you are, or know of someone who may be interested please contact the
director of the lab listed below
Simon C. Watkins
Director SBIC
840 Scaife Hall
University of Pittsburgh
Pittsburgh PA
412-648-3051
email Swatkins-at-Pitt.edu






From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Tue, 8 Nov 1994 08:07:59 -0600
Subject: Re: Spelling errors / typos

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Bill,

Its alright, Bill. In the spirit of academic freedom & creativity I'm perfectly
willing to accept your spelling of "Copland". Or to put it to you as a quote
from our oatmeal poet-lorryate[sic], Bob Dylan, "Don't think twice, its
alright......Babe".



In message {9411072352.AA03570-at-MOLE.mmm.com} writes:
} One last (?) word on spelling errors:
} "It's a damn small mind that can think of only one way to spell a word."
} - Andrew Jackson (seventh President of the U.S.)

Gib

P.S. - to follow soon. Must go aline the CM12 for 120kV.





From: Katherine.S.Connolly-at-Dartmouth.EDU (Katherine S. Connolly)
Date: 08 Nov 94 10:05:36 EST
Subject: Recoating phosphor screens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {8593967-at-dancer.Dartmouth.EDU}

Some years ago Ladd Industries, PO Box 1005, Burlington Vt. 05402, did a fine
job recoating a screen for my Philips TEM.
Kate




From: Ronald Henry Birkhahn :      rbirkhah-at-uceng.uc.EDU
Date: Tue, 08 Nov 1994 09:55:18 -0500 (EST)
Subject: Re: Tiff images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree with Scott, Adobe Photoshop is the best way to view/manipulate
images, but for the Mac, there is a shareware program called GIFconverter
2.3.2 which will open any picture file and convert to any known format you
can think of.

Ron




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Tue, 8 Nov 1994 11:57:42 -0500 (EST)
Subject: RE: spelling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


"You can't help respecting anybody who can spell TUESDAY, even if thy
don't spell it right; but spelling isn't everything. There are days when
spelling Tuesday simply doesn't count." A.A. Milne. from The House at
Pooh Corner.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************





From: Ronald Henry Birkhahn :      rbirkhah-at-uceng.uc.EDU
Date: Tue, 08 Nov 1994 09:55:18 -0500 (EST)
Subject: Re: Tiff images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1427819117.53101-at-mse.engin.umich.edu}

John F. Mansfield
North Campus Electron Microbeam Analysis Lab.
2455 Hayward Ann Arbor MI 48109-2143
Tel: (313)936-3352
Email: jfmjfm-at-engin.umich.edu or John.F.Mansfield-at-umich.edu
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html
Time:
12:59
Date:
11/8/94

Graphic Converter is better & supports more formats.

--------------------------------------

Ron

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;8 Nov 1994 12:20:25 -0400
Received: by judgmentday.rs.itd.umich.edu (8.6.9/2.2)
with X.500 id LAA22095; Tue, 8 Nov 1994 11:18:29 -0500
Received: from AAEM.AMC.ANL.GOV by judgmentday.rs.itd.umich.edu (8.6.9/2.2)
with SMTP id LAA22076; Tue, 8 Nov 1994 11:18:23 -0500







From: Richard E. Edelmann :      REDELMAN-at-musom01.mu.wvnet.edu
Date: Tue, 08 Nov 1994 13:24:11 +1100
Subject: PC Operating Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I am looking for feed back from anyone out there listening.

I am wondering how UNIX, Windows NT and OS/2 compare overall but
especialy how the latter two compare with UNIX. I am looking into
upgrading operating systems. In general the hardware platform will
be a 586/90MHz PCI/EISA Bus, 16 Meg RAM, 1 GB diskspace, eithernet,
etc.. I would like to go with UNIX (Novell or SCO), but I am
concerned about possible problems with running the vast array of
software written for the windows world in the X-windows environment
and if the windows software would run better under Windows NT (Or
OS/2).

Does anyone out there have any experiences to share? General
preferences? Considerations or comments?

If you E-mail directly I'll try and coallate the info and either
post it as one message or make it available to anyone who'd like it.

Thanks in advance.


Richard E. Edelmann
Electron Microscopy Facility Supervisor
Marshall University - School of Medicine
Huntington, West Virginia




From: {John_Mansfield-at-mse.engin.umich.edu}:ddn:wpafb
Date: 11-8-94 1:01pm
Subject: RE>> format converters, was RE>Tiff images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411082037.AA25076-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: Re: format converters, was
Orig-Author: {"John Mansfield" {John_Mansfield-at-mse.engin.umich.edu} }:ddn:wpafb
-----------------------------------------------------------
John F. Mansfield
North Campus Electron Microbeam Analysis Lab.
2455 Hayward Ann Arbor MI 48109-2143
Tel: (313)936-3352
Email: jfmjfm-at-engin.umich.edu or John.F.Mansfield-at-umich.edu
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html
Time:
12:59
Date:
11/8/94

Graphic Converter is better & supports more formats.






From: Dave DeFily :      defily-at-tam2000.tamu.edu
Date: Tue, 8 Nov 1994 15:29:22 -0600
Subject: No Subject

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


suscribe defily-at-tamu.edu

-Dave defily-at-tamu.edu





From: tony-at-emu.su.oz.au (Tony Romeo)
Date: Wed, 9 Nov 1994 07:43:31 +0100
Subject: Re: recoating phosphor screens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411082137.AA24911-at-ELECTRON.emu.su.OZ.AU}
X-Sender: tony-at-electron.emu.su.oz.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear Peter,
I actually run a screen recoating service from the E.M. Unit
here at very reasonable rates. If you are interested in any further
information contact me direct.

Cheers,

Tony Romeo

Tony Romeo Internet: tony-at-emu.su.oz.au
Electron Microscope Unit
Madsen Building F09 Telephone: +61 2 351 2351
The University of Sydney Facsimile: +61 2 552 1967
Sydney, NSW 2006
Australia







From: Arthur_Strange-at-mail.magic.ca
Date: Tue, 08 Nov 1994 16:39:26 EST
Subject: Re: Tiff images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {1994Nov08.163926.3183459-at-mail.magic.ca}
To: microscopy-at-aaem.amc.anl.gov

Hey Ron, I think you have the programmes mixed up, GraphicConverter v2.0 by
Thorsten Lemke converts most any graphic file. In addition you can do a
little image processing.

Arthur





From: mwfolsom-at-hydra.unm.edu (Mike Folsom)
Date: Tue, 8 Nov 1994 15:11:55 -0700 (MST)
Subject: embedding media & thick sections .....

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0r4ylg-0009jZC-at-hydra.unm.edu}

Folks -

A quick question -

I'm trying to embed plant material is some type of matrix that will
allow me to make fairly thick sections, say 25 - 50 um.

I've used paraffin and the tissue starts to fracture when section
thickness gets greater than 20 - 25 um. Frankly besides embedding
the tissue in plastic and then grinding it down I'm not sure
what else to do.

I'd appreciate any suggestions -

Michael

_______________________________________________________________________________
M.W.Folsom/Biology/UNM/Albuquerque,NM~87131/505.277.4277/mwfolsom-at-hydra.unm.edu





From: tvoiles-at-unlinfo.unl.edu (todd voiles)
Date: Tue, 8 Nov 1994 17:00:13 -0600 (CST)
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe broberston-at-unlinfo.unl.edu

--

E.Todd Voiles *** Facility Manager
Central Facility for Electron Microscopy
University of Nebraska
tvoiles-at-unlinfo.unl.edu




From: Ronald Henry Birkhahn :      rbirkhah-at-uceng.uc.EDU
Date: Tue, 08 Nov 1994 17:13:33 -0500 (EST)
Subject: Re: format converters, was

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bob,
I don't know about Graphic Converter that John was talking about, but
most images that are passed around these days (that I know of) are either
in TIFF, GIF, or JPEG format. This program will convert between those
and a a handful of others. I think i pulled it off of an Apple shareware
site. I think the U. of Texas has one. If you can't find it, I could
try sending it to you compacted. But check out the one Jahn Mansfield
was talking about first.

Ron





From: Colin Veitch :      VEI011-at-GEEL.DWT.CSIRO.AU
Date: Wed, 9 Nov 1994 8:52:46 +1100 (EST)
Subject: Re: Tiff Image Viewers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We use Adobe Photoshop (both PC and Mac) and have found it very powerful
and useful. It is however expensive (at least in Australia). JASC
produce a Windows product called Paint Shop Pro which is not as
expensive (around $70 USD) but can read and write many (around 40)
image formats including most of the more common TIFF formats. It also has
contrast/brightness controls, gamma correction and some filters.

I think, though I'm not sure, that there is a shareware version
available.

According to the manual the address and 'phone number is:

JASC Inc.
10901 Red Circle Drive, Suite 340
Minnetonka, MN 55343
(612) 930-9171

I have no interest in the company, I just use the software!!

Other programs which may be of use are Image Alchemy (PC) or NIH
Image (Mac).

Good Luck!!


#####################################################################
**********************
* Between the idea *
0------* And the reality *
} ---|--- { * Between the motion *
| * And the act *
/ \ * Falls the Shadow *
_/ \_ * T.S. Eliot *
**********************
Colin Veitch Tel + 61 (0)52 27 5611
CSIRO Division of Wool Technology Tel + 61 (0)52 27 5891 (dir.)
P.O. Box 21 Fax + 61 (0)52 27 5657
BELMONT Vic 3216
Australia

#####################################################################




From: Fred Schamber :      73751.3677-at-compuserve.com
Date: 08 Nov 94 17:15:46 EST
Subject: Magnetic shielding principles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There have recently been several items posted concerning magnetic shielding of
EM rooms. Several people have pointed out that this is an expensive and
generally impractical thing to attempt, and they are right.

Something that is often not appreciated is that you don't actually "shield"
magnetic fields in the sense of "blocking" them. Instead you "shunt" the fields
around the area you are trying to protect. The commonly used shielding
materials have extremely high initial permeability which means that magnetic
fields will travel through them in preference to other routes. In other words,
you don't build a "dam" in the way of the field, but instead you provide a
"sponge" to channel it around the area to be shielded.

This sounds fairly simple until you realize that the high permeability of the
material means that you can also easily channel magnetic fields INTO the area
you are trying to shield if the enclosure is not a properly designed magnetic
circuit. Just sticking a sheet of "mu-metal" in the way of the field is likely
to make things worse. Holes or open joints can also create "lensing" effects.

Even in the best of cases, all of the magnetic flux won't flow through the path
provided by the "shield" -- some amount of field always fringes out into the
surrounding area. So at best, all you can do is attentuate the fields (not
eliminate them entirely). The more efficient the circuit you provide for the
fields to travel in, the greater the attenuation. However, as enclosures
increase in size, more of the field lines will find it advantageous to take a
"short cut" across the enclosed free-space region. For a simple cylindrical
field, the attenuation varies as the permeability times thickness of the shield,
divided by its diameter. In practical terms, this says it's pretty easy to
shield a small region, but to get comparable shielding of a room, for example,
the shield has to be quite thick (or consist of multiple isolated layers).
Obviously, it is best to shield the smallest possible volume, which is why you
want the shielding designed into your EM, rather than applied as an
afterthought.

By way of reference, the initial permeability of a typical 80/20 (Ni/Fe)
shielding material will be in excess of 20,000, as compared to several hundred
for "soft" iron (free space is 1.0). Oriented silicon steel, as used in
transformers, has IP of about 1000 and is often a good shielding material.

The above remarks apply principally to the case of lower frequency fields (such
as the omnipresent 60 Hz). For high frequency fields one gets substantial
eddy-current shielding just by using a good electrically conductive enclosure,
such as aluminum.

There is real art to designing a good shield, and for all but the simplest
cases, one is well advised to seek the assistance of someone who does it
routinely. I'm not such a person, but have had good luck in the past working
with Advance Magnetics (219) 223-3158, whose catalog also contains a lot of
helpful application information (I have no connection to this firm). There are
also a number of other good companies in this business.

By the way, I have been looking for years for an inexpensive PC-based program
which would be useful for designing magnetic shields -- ideally something in the
freeware/shareware realm. If someone out there knows of such a thing (or knows
of someone who might), I'd appreciate hearing about it.

Fred Schamber
RJ Lee Group, Instruments Div.







From: Doug Arrell :      ARRELL-at-jrc.nl
Date: Wed, 9 Nov 1994 08:24:44 GMT+0200
Subject: Re: Tiff images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.941109082444.256-at-FS-IAM-1.JRC.NL}

Another good shareware alternative for viewing TIFF, or just about
anything else, is Paintshop Pro. It claims to handle three or so
types of compression for TIFF files, up to 32 bit colour depth etc. I
don't know whether it will cope with _all_ TIFF files but it works
for everything I have tried it with.

Good Luck

Doug

+------------------------------------+
| Dr Douglas Arrell |
| Mechanical Performance and Joining |
| Institute for Advanced Materials |
| 1755 ZG Petten |
| Netherlands |
| {ARRELL-at-JRC.NL} |
| Tel. (+31) 2246 5287 |
| Fax (+31) 2246 1917 |
+------------------------------------+




From: Richard E. Edelmann :      REDELMAN-at-musom01.mu.wvnet.edu
Date: Wed, 09 Nov 1994 08:05:54 +1100
Subject: RE: PC OPerating Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have been questioned as to the validity of my inquiry of PC
Operating systems on this BBS, let me explain.

I am setting up a local network system to intergrate my EM
Facility. This platform will be the backbone server for the exchange
of data (Images from LM, LSM, TEM, SEM, EDS, etc), such that users
can manipulate their images at a variety of sites. It is being
designed to be networked directly to microscopes which use DOS,
Windows, and X-Windows. This net will also allow for sellection
between a variety of hardcopy output devices. It is also the
intention that this system serve as an internet node allowing for the
exchange of images across the net.

The platfrom will also be used for tasks such as E-mail,
word processing, statistical analysis.

The microscope/networking end is covered. However I was
interested in collecting further info for the operating system to
function under. Since the users of this BBS are very familar with
computer systems and complex software (i.e. high end imaging
software) I thought this was a good place for this question.

I am sorry if you disagree.

Yes, this type of a system could be purchased from a vendor out
there for however many $10k's (and for those of you who can afford it
great) but the I know many labs out there are like mine in which we
need to do everthing we can to save cost, and still get the best we
can afford.


Richard E. Edelmann
Electron Microscopy Facility Supervisor
Marshall University - School of Medicine
Huntington, West Virginia




From: ldm-at-apollo.numis.nwu.edu (L. D. Marks)
Date: Wed, 9 Nov 1994 08:19:49 -0600
Subject: Magnetic shielding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A number of people have commented recently on how difficult it
is to shield stray fields, suggesting that one needs to shield a whole
room at the cost of } $10K. I would like to point out that simple, and
cheap shielding with a little Mumetal has, EXPERIMENTALY, worked in at
least two cases I know of:
a) The old Cambridge HREM where a small shield was introduced in
the column. (Ask Dave Smith now at ASU for further details.)
b) Our own HREM here at Northwestern.

True, such a simple shield may not completely cure the problem,
but for a few hundred dollars it is well worth the attempt! A very simple
test that my students devised was to wave a small magnet near the
microscope to test what regions were sensitive.

I should also comment that most people build small (cheap) shields
around LEED systems, and most surface science systems operating with low
energy electrons use shields with holes in them (for flanges). The VG
STEMs also have a shield above the objective which has holes in it.

L. D. Marks
Northwestern




From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Wed, 9 Nov 1994 09:53:28 -0500
Subject: Re: Recoating phosphor screens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



On 8 Nov 1994 RETEP-at-anat.uct.ac.za wrote:

} Dear All,
}
} Anyone out their re-coat their phosphor screens?
}
} How does one go about it? Are the supplies available in South Africa?
}
} The screen I am proposing to coat has an alluminium base, about 12 cm
} diameter, for a Hitachi TEM.
}
} Many thanks for any advice.
}
} Peter Richards.
}
} _______________________________________________________________
}
}
} -at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
} -at- -at-
} -at- Peter D. G. Richards -at-
} -at- Dept Anatomy and Cell Biology -at-
} -at- UCT Medical School -at-
} -at- Observatory -at-
} -at- 7925 -at-
} -at- RSA -at-
} -at- Tel: 021-406 6285. -at-
} -at- Internet: retep-at-anat.uct.ac.za -at-
} -at- -at-
} -at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
} Peter:
Recoating screens is a fairly simple thing to do. I used to recoat mine
all the time using a technique used by a now defunct company called AEI.
I used to use a phosphor put out by JEOL. Always got beautiful screens.
Their phosphor gave me a screen with good contrast levels. If you can
give me a fax number, I will fax a copy of the technique to you. I still
have some of the phosphor and could send some to you if there isn't any
problem with the mail system getting it to you.
Let me know.
Phil





From: Fred Schamber :      73751.3677-at-compuserve.com
Date: 09 Nov 94 11:13:18 EST
Subject: TIFF Image Converters

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is to add on to the message posted by Colin Veitch:
} } JASC produces a Windows product called Paint Shop Pro which is not as
expensive (around $70 USD) but can read and write many (around 40)
image formats including most of the more common TIFF formats. { {

This is a very nice product. It is shareware and can be downloaded from
Compuserve, AOL, etc. The registration fee is $69. It lists at least 20
discrete formats (GIF,TIF,JPG, etc) and seems to handle them all smoothly,
including converting between.

JASC also sells a companion Windows product called "Image Commander" which
allows you to setup "catalogs" of images and view them as a page of thumbnail
pictures or select a particular one for viewing. It has provisions for
organizing the catalogs, annotating the images, etc., and would look like a very
useful thing for anyone handling lots of digitized micrographs. This is also
shareware, and the registration is $39.

I have no interest in the company, but have had good experience with both
products.

Fred Schamber
RJ Lee Group, Instruments Div.





From: tvoiles-at-unlinfo.unl.edu (todd voiles)
Date: Wed, 9 Nov 1994 10:48:12 -0600 (CST)
Subject: MSDS's Online

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Recently I saw a posting about an online source for MSDS's but
have now mislaid the info....could someone help me out for
safety's sake?

Thankx

--

E.Todd Voiles *** Facility Manager
Central Facility for Electron Microscopy
University of Nebraska
tvoiles-at-unlinfo.unl.edu




From: Arthur_Strange-at-mail.magic.ca
Date: Wed, 09 Nov 1994 13:47:25 EST
Subject: Re: TIFF Images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {1994Nov09.134725.3200862-at-mail.magic.ca}
To: microscopy-at-anlemc.msd.anl.gov

S

Hey Ron, I think you have the programmes mixed up, GraphicConverter v2.0
by
Thorsten Lemke converts most any graphic file. In addition you can do a
little image processing.

Arthur





From: NANCY SMITH :      NSMITH-at-darwin.sci.csuhayward.edu
Date: Wed, 9 Nov 1994 09:42:35 PSD8PDT
Subject: pc operating systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I am gathering and storing away much of the information on this
bulletin board regarding networking of microscopes and image storage
systems. I, for one, appreciate the inquiries of people such as
Richard Edelmann regarding computers and operating systems for
microscopy imaging. Also, I have noticed in the last week that my
spelling has improved vastely.

Nancy Smith
Director, EM Lab
CSU,Hayward
Hayward, CA 94542




From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Wed, 9 Nov 1994 12:57:12 -0500 (EST)
Subject: RE: PC OPerating Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Although you may end up using it inelegantly, e.g. via telnet and ftp,
probably the way to go is:
o Unix machine with tons of disk space as a server when images are stored
temporarily
o stand alone PCs (or Macs, etc.) have their own removable media
and users MUST purchase their own disks for storage.

We have found that attempting to network PCs (some DOS, some windows, some
OS2), UNIX, OS7, and Macs is nearly impossible. The only platform that
appears to be compatible with all the other platforms is UNIX.

-MC






From: EMLAB-at-opus.mco.edu
Date: Wed, 09 Nov 1994 15:44:59 -0400 (EDT)
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Michael,

Try embedding in 6% agarose, then vibrotoming(sp?).

Ed Calomeni




From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 9 Nov 1994 18:14:01 -0400
Subject: Re-Spelling Fluorine

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1427717408.73193-at-mse.engin.umich.edu}

Subject: Time:5:05 AM
OFFICE MEMO Re:Spelling Fluorine Date:11/10/94

Like the professor who pointed out that if the first five letters
spelled 'flour' it was wrong, I had a biochemistry professor who
told us to split the word into two syllables, and if it came out
'flo-urine', it was wrong.





From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Wed, 9 Nov 1994 14:25:59 -0700
Subject: TEM available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411092125.AA136454-at-lamar.ColoState.EDU}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Our department has available, for the cost of moving, a Hitachi HS-8 TEM.
It is a one owner instrument and has been maintained on manufacturer's
service contract since installation. It is currently fully operational.
We will be moving this microscope very soon.

Please contact me directly for more information.


John chandler-at-lamar.ColoState.EDU Fort Collins CO






From: Kimberly Wilson :      kwilson-at-admin.ogi.edu
Date: Wed, 9 Nov 1994 09:40:07 -0800 (PST)
Subject: questions about suppliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I recently spend 4 weeks in the Netherlands learning microscopy for
yeast samples-only morphology and immunogold techniques. Now I am
trying to set up our lab in America to do these techniques. I have
three major problems-limited experience, no one at my institute can
help me, and having learned in Europe I have no knowledge of good
American suppliers for EM stuff.

Here is what I need:
Right now we have a sorvall MT2-B "Porter-Blum" ultra-microtome.
We are looking for a table for it (vibration resistant), as well as
for a new ultramicrotome.

I also would like to know a good general supplier (other than EM sciences)
for things like grids, grid holders, capsule holders, etc....

I also would like to know where people get good quality formaldehyde,
glutaraldehyde, potassium permanganate, and ethanol ( in other words
the basic chemicals needed).

Thanks for your help.
Kim Wilson-Oregon Graduate Institute

kwilson-at-admin.ogi.edu




From: tivol-at-tethys.ph.albany.edu
Date: Wed, 09 Nov 1994 10:31:13 EST
Subject: Re: Recoating screens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In article {MAILQUEUE-101.941108141914.288-at-anat.uct.ac.za} ,
RETEP-at-anat.uct.ac.za writes:
} Dear All,
} Anyone out their re-coat their phosphor screens?
} How does one go about it? Are the supplies available in South Africa?
} The screen I am proposing to coat has an alluminium base, about 12 cm
} diameter, for a Hitachi TEM.
}
} Many thanks for any advice.
}
} Peter Richards.

Dear Peter,
We routinely recoat our screens, and, if the phosphor is available, the
rest of the supplies will be easy. The steps are: 1) Make } 500 ml 0.5% gel-
atin solution, 2) Filter with type GS 0.22 mu millipore -at- ~30 C, 3) Suspend
10.6 g (for ~45 g/cm**2 and ~170 cm**2 screen size, your mileage may vary)
phosphor in 500 ml gelatin soln, 4) Warm and degas the suspension while, 5)
Clean screen blank, 6) Glow discharge blank for 15 min, 7) Place blank in
funnel so that it is horizontal and so that there is an area between the edge
of the blank and the wall of the funnel (our screen blank is not completely
circular, but has two straight sides cut along chords, you may have to build a
frame to achieve this geometry), 8) Attach a hose to the bottom of the funnel,
and clamp it off, 9) Fill the funnel with 50% ethylene glycol to a level just
above the bottom face of the blank and make sure there are no air bubbles trap-
ped beneath the surface, 10) Pour the phosphor suspension down a glass rod and
move the rod around to avoid pileup on any one area, 11) Cast a film of collo-
dion diluted 1:100 in n-butyl acetate on the surface of the liquid by deposi-
ting drops with a Pasteur pipette until the entire surface is covered, 12) When
the n-butyl acetate has evaporated, sweep the film aside with a wedge of filter
paper; repeat until the liquid surface is free of debris, 13) open the clamp on
the hose and allow the liquid to drain slowly--~2 drops/sec--until the level is
below the bottom surface of the blank, 14) Place a clean sheet of paper over
the top of the funnel and draw air through the hose and past the screen; leave
overnight, and 14) Remove the screen from the funnel, check for evenness of
coating and lack of foreign objects (eyebrows etc.), then bake at 60 C for 24
hours. Good luck, it takes a little practise, but the blanks can, of course,
be reused.
Yours,
Bill Tivol




From: KAKER-at-ctklj.ctk.si
Date: Thu, 10 Nov 1994 8:22:45 +0100 (WET)
Subject: RE:EM suppliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


1. Structure Probe Inc.
SPI Supplies
P.O.Box 656
West Chester,PA 19381-0656

Tel:215-436-5400
Fax:215-436-5755

girds,SEM mounts,chemicals,diamond knivesµtomy,standards,
small tools,instruments for specimen preparation

2. Ernest F.Fullam Inc.
900 Albany Shaker Road
Latham,NY 12110

Tel:800-833-4024
Fax:518-785-8647

3. Ted Pella Inc.
P.O.Box 2318
Redding,CA 96099

Tel:916-243-2200
800-637-3526 (CA)
Fax:916-243-3761

4. BAL-TEC Products Inc.
984 Southford Road
Middlebury,CT 06762

Tel:203-598-3160
Fax:203-598-3658


Henrik Kaker




From: lebail-at-cnrs-imn.fr
Date: Thu, 10 Nov 1994 11:48:37 GMT
Subject: TEM and inorganic fluorides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Fluorine being on the scene, I take my chance once more to ask again
a question stayed without any response since two months :

"How to avoid the almost systematic decomposition of inorganic fluoride
compounds under the electron beam during a TEM experiment (JEOL 2010) ?
Any suggestion appreciated (inversely as the cost)."

The question may be formulated in another way : Has somebody ever worked
with some success on inorganic fluorides by TEM ?

An element of response is in "The electron microscopic observation of
phase transition in KFeF4, RbFeF4, and RbVF4," by R. Deblieck, J. Van
Landuyt & S. Amelinckx, J. Solid State Chem. 59, 379-387 (1985). The
problem with inorganic fluorides is electrical conduction :
"After washing in benzene, in order to remove tape residues, the
thinned plates are mounted on 3-mm copper grids by means of silver
paste to ensure good thermal and electrical contact. In spite of
these precautions electrical conduction remains a problem and
substantial charging phenomena still occur under the electron
bombardment for observation. Moreover, all materials are subject
to damage by electron irradiation..... The investigated compounds
prove to be sensitive to decomposition under electron irradiation,
probably through temperature-driven evaporation of fluorine."

Other experiences on inorganic fluorides by TEM ? Suggestions ?

Microscopists do not like fluorine, this may be why there are some
spelling problems.

Armel Le Bail - Fluoride Lab - armel-at-ONE.univ-lemans.fr




From: {lebail-at-cnrs-imn.fr}:ddn:wpafb
Date: 11-10-94 7:26am
Subject: Re: TEM and inorganic fluorides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411101344.AA06245-at-riker.ml.wpafb.af.mil}
To: WALCKSD:ml:wpafb, SCHELTFJ:ml:wpafb
Subj: TEM and inorganic fluorides
Orig-Author: {lebail-at-cnrs-imn.fr}:ddn:wpafb
-----------------------------------------------------------
Hi,

Fluorine being on the scene, I take my chance once more to ask again
a question stayed without any response since two months :

"How to avoid the almost systematic decomposition of inorganic fluoride
compounds under the electron beam during a TEM experiment (JEOL 2010) ?
Any suggestion appreciated (inversely as the cost)."

The question may be formulated in another way : Has somebody ever worked
with some success on inorganic fluorides by TEM ?

An element of response is in "The electron microscopic observation of
phase transition in KFeF4, RbFeF4, and RbVF4," by R. Deblieck, J. Van
Landuyt & S. Amelinckx, J. Solid State Chem. 59, 379-387 (1985). The
problem with inorganic fluorides is electrical conduction :
"After washing in benzene, in order to remove tape residues, the
thinned plates are mounted on 3-mm copper grids by means of silver
paste to ensure good thermal and electrical contact. In spite of
these precautions electrical conduction remains a problem and
substantial charging phenomena still occur under the electron
bombardment for observation. Moreover, all materials are subject
to damage by electron irradiation..... The investigated compounds
prove to be sensitive to decomposition under electron irradiation,
probably through temperature-driven evaporation of fluorine."

Other experiences on inorganic fluorides by TEM ? Suggestions ?

Microscopists do not like fluorine, this may be why there are some
spelling problems.

Armel Le Bail - Fluoride Lab - armel-at-ONE.univ-lemans.





From: tvoiles-at-unlinfo.unl.edu (todd voiles)
Date: Thu, 10 Nov 1994 09:47:56 -0600 (CST)
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe brobertson-at-unl.edu

--

E.Todd Voiles *** Facility Manager
Central Facility for Electron Microscopy
University of Nebraska
tvoiles-at-unlinfo.unl.edu




From: Carl Henderson :      chender-at-umich.edu
Date: Thu, 10 Nov 1994 11:09:23 -0500 (EST)
Subject: Re: Laser printer enhancements?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Steve,

If necessary, you can add additional NuBus cards to your Mac by adding a
NuBus expansion chassis. I have one made by Second Wave which gives me an
additional three slots (as well as a power supply and connections for an
internal hard disk drive) on my Mac IIci. There are also six slot chasses
available.

Second Wave can be contacted at:
9430 Research Blvd., Echelon II, Suite 260
Austin, TX 78759-6541
512-343-9661
512-343-9663 (FAX)
Applelink: D0864

I have no connection to Second Wave, I'm just a satisfied customer.

Carl Henderson
Electron Microbeam Analysis Laboratory
University of Michigan
2005 C.C. Little Bldg.
Ann Arbor, MI 48109-1063 USA
-----------------------------------------------------
(313) 936-1550 (voice) **** Next time, ****
(313) 763-4690 (FAX) *** take the ***
chender-at-umich.edu (e-mail) **** train! ****
-----------------------------------------------------
On Thu, 10 Nov 1994, Steve Bill 5355/5128/5642 wrote:

}
} We are currently using a HP Laserjet 4M to output TEM images from our Gatan
} Slow-scan camera (mac based). What I'm looking for is an enhancement that will
} improve the output of images from this printer. I've heard of some grey-scale
} sofware upgrades and some add-in cards that are supposed to boost the resolution
} or improve the grey-scale rendition of the HP laser printers but I don't have any
} specific information or pointers to any of them.
}
} The Mac by the way has no NUBUS slots available (all three are taken) so anything
} we consider must either be software-based or hardware added to the printer itself.
}
} Thanks,
}
} Steve Bill
}




From: tivol-at-tethys.ph.albany.edu
Date: Thu, 10 Nov 1994 12:21:02 EST
Subject: TEM and inorganic fluorides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411101659.AA16230-at-MIT.EDU}

Dear Armel et al.,
My experience with fluorides is limited to organics, but the problem
of radiation damage is still serious with these fluorides as well. The best
procedure IMHO is to use as low a dose as possible--especially for focusing,
etc., where the electrons are causing damage, but not being recorded. We
use our fairly crude, but sensitive, video system to scan grids for areas of
interest, focusing, etc., and LoDose or SO163 for image recording.
I have found that if the lens column is aligned well, the wobbler can
be used to focus quickly within one click of the fine objective control (=.99
micron). A subsequent through-focus series should yield good images. This
works at mags up to 200,000, and for low dose images, I have had my best luck
with somewhat lower mags than this.
I use a 30 micrometer C2 aperture with the wehnelt bias on the lowest
setting and the C1 lens maximally excited. This produces a beam current in
the picoamp range, where charging, heating, etc. are minimized (it helps to use
1.2 MV electrons if you have them).
Remember, alignment of the lens column is critical--check and correct
it the same day you take the pictures. Good luck.
Yours,
Bill Tivol




From: jacobb-at-ux5.lbl.gov
Date: Thu, 10 Nov 1994 09:42:53 -0800
Subject: cryo EM of fluorides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to: "How to avoid the almost systematic decomposition of
inorganic fluoride compounds under the electron beam during a TEM
experiment...?"

You might try cryo EM, that is lowering the temperature of the specimen
towards that of liquid nitrogen or liquid helium. This preserves visible
structure for a time in biological specimens, presumably due to a decreased
vapor pressure of the products of radiolysis.

We are struggling with imaging droplets of liquid fluorinated hydrocarbons,
by freezing the drops on a substrate and viewing them with the
low-temperature scanning electron microscope at -180 C. At low
temperature, large (1mm diameter) drops are resistant to the 10kV beam,
small (1 m) droplets sublime. Interestingly, even after a ca. 100A gold
coat the fluorocarbon drops appear brighter than adjacent aqueous
structures, perhaps due to charging or the differing secondard electron
coefficients of fluorocarbon and water.

Jacob Bastacky, MD
1-116
Lawrence Berkeley Laboratory EMail: sjbastacky-at-lbl.gov
University of California Phone: (510) 486-4606
Berkeley, California 94720 Fax: (510) 486-4750





From: South Bay Technology :      73531.1344-at-compuserve.com
Date: 10 Nov 94 12:57:56 EST
Subject: questions about suppliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would suggest that you contact one of the following:

SPI Supplies
800-242-4774

Ernest F. Fullam, Inc.
800-833-4024

Ted Pella, Inc.
800-237-3526

Bal-Tec Products, Inc.
800-875-3713

Good luck-

David Henriks
South Bay Technology, Inc.
1120 Via Callejon
San Clemente, CA 92673

TEL: 714-492-2600
FAX: 714-492-1499





From: Greg Begin :      GREGB-at-sales.lmt.com
Date: Thu, 10 Nov 1994 11:46:30 CST6CDT
Subject: Re: Laser printer enhancements?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Steve Bill 5355/5128/5642 {sbill-at-asdg.enet.dec.com}

Steve,

Information about LaserPrinter Enhancements:

PC based controllers are available from the following companies;

LaserMaster Corporation - WinJet Product Line (Windows Only)
Talltree Systems- JLaser? Greylaser? (? if still selling)
DPTech - ? (not sure if they are still selling units)
XLI Corp - ?(not sure of Name) Should be selling controllers only No
complete products including engine.

Mac Based products; I am currently not aware of any in a controller
or software only basis.

Complete Printer subsystem for High-Res output -- PC & MAC & Unix*

LaserMaster ImagePrinter 1800 DPI (#2 Version)
Appletalk, Eithertalk, TCP/IP, Novel Netware, Parallel (HS),
Serial communications on unit.
24 meg Ram
66 Mhz Multi-Tasking Pipeline Processor
240 Meg Hard Drive internal
Pipeline Assoc. Postscript Level 2 interpreter & PCL 4 Languages
8 1/2 X 11" to 12.5 X 19" paper size on line.

*unix is only for TCP/IP standards on limited System OS

This is only 1 of our printers avaiable. I do work for LM in the
Scientific and Medical industries and can help you if you would like.
Contact Greg Begin LM Imaging at 1-800-950-6363 Ext: 3207 or Mail
address of gregb-at-sales.LMT.com I will be out at the American Heart
Assoc. meeting next week and then the Radiological Society of North
American durring the last week of Nov.
Greg Begin LM Imaging Div. LaserMaster Corporation
/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\

} Date sent: Thu, 10 Nov 94 09:59:20 EST
} From: Steve Bill 5355/5128/5642 {sbill-at-asdg.enet.dec.com}
} To: microscopy-at-aaem.amc.anl.gov
} Copies to: sbill-at-asdg.enet.dec.com
} Subject: Laser printer enhancements?

}
} We are currently using a HP Laserjet 4M to output TEM images from our Gatan
} Slow-scan camera (mac based). What I'm looking for is an enhancement that will
} improve the output of images from this printer. I've heard of some grey-scale
} sofware upgrades and some add-in cards that are supposed to boost the resolution
} or improve the grey-scale rendition of the HP laser printers but I don't have any
} specific information or pointers to any of them.
}
} The Mac by the way has no NUBUS slots available (all three are taken) so anything
} we consider must either be software-based or hardware added to the printer itself.
}
} Thanks,
}
} Steve Bill
}




From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Thu, 10 Nov 1994 16:12:50 -0500
Subject: Coating screens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


After so many requests for the procedure, it is a little cheaper to put
it out on the microscopy board. So, here goes for those who want it.
I get my phosphor from JEOL, cat.#423-011(green) or 423-010(yellow).
These are old cat. numbers, they might have changed. You can call the
parts dept. at (508) 536-2342 and ask for Maria Ramos.

1. Withdraw the old screen from the microscope, taking care to avoid
depositing any loosened phosphor in the viewing chamber.
2. Remove the old coating from the screen plate by washing in acetone.
The cleaned plate must be free from all particles of matter and the
surface must be free from blemishes and scratches.
3. Prepare a sufficient volume of 4% (w/v) suspension of phosphor powder
in acetone containing about 1% of collodion. To make collodion use:
4grams parlodion, 25ml 100% ETOH, 75ml ether.,
Any grade of collodion that is suitable for specimen preparation can be used.
The total volume must be sufficient to fill the selected dish with liquid
to a depth of about 1cm above the surface of the screen plate in position
on the bottom of the dish.
4. Agitate the suspension vigorously then pour it rapidly into the dish.
5. Wait about 5 seconds to allow large particles to settle and violent
swirling to cease.
6. Slide the screen smoothly into the liquid, preferably without
scraping the bottom of the dish.
7. Cover the dish with a piece of plexiglass with a small (1/4"-1/2")
hole drilled in it and leave it to settle (I usually let it sit overnight).
When the remaing liquid is clear, draw off the liquid by inserting a
suction tube through the hole in the lid. It is extremely important not
to disturb the screen or the liquid above the screen in any way as this
is being done. Draw off the liquid steadly and slowly then remove the
suction tube.
8. Leave the screen to dry without any further disturbance of any kind.
Do not lift the lid to inspect the screen until the powder is quite dry
because a slight change in drying conditions can produce a visible mark
on the damp surface.
9. When the screen is dry, remove the screen and wipe off any excess
phosphor from the back and sides of the screen plate.
10. Put screen back into the microscope.

Note: A newly-coated screen will de-gas for a short time when it is
first placed in the microscope. Pumping times may therefore be longer
than normal at first.

A very small particle size is desirable for high resolution screens and
it may be found advantageous to agitate the phosphor suspension
ultrasonically before putting it in the dish. Screen resolution is also
affected by the thickness of the phosphor layer and by light scattering
within it.

It may take a few tries to get the hang of it, but when I had to recoat
my screens this technique worked rather well. Fortunately, I no longer
have to recoat my screens.

Good luck in trying this!
Phil




From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Thu, 10 Nov 1994 18:14:47 -0400 (EDT)
Subject: RE: Waste OsO4

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

According to a OSHA publication (1978), liquid containing Osmium tetroxide
should be absorbed in vermiculite, dry sand or similar material and then
disposed off in a sealed container in a secured sanitary landfill.




From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Thu, 10 Nov 1994 17:01:22 -0600
Subject: Re Os disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
There was an interesting paper in the Recent Microscopy
Today (october) in which different ways of neutralizing liquid osmium
waste/spills are tested. They find corn oil-soaked kitty litter to be the
best treatment of all, with pure corn oil not far behind.

Tobias Baskin

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
___ ____ ^ ____ _____ Tobias I. Baskin
/ \ / / \ / \ / University of Missouri
/ | / / \ / / Biological Sciences
/___ / /__ /_____\ / /__ 109 Tucker Hall
/ / / \ ( / Columbia, MO 65211 USA
/ / / \ \ / voice: 314-882-0173
/ /____ / \ \____/ /_____ fax: 314-882-0123






From: Eero Ristolainen :      ert-at-leka.hut.fi
Date: Fri, 11 Nov 1994 09:09:11 +0200 (GMT)
Subject: Returned mail: Host unknown (Name server: anlemc.msd.gov: host not found) (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Forwarded message:
From daemon Thu Nov 10 14:19:48 1994



From: Eero Ristolainen :      ert-at-leka.hut.fi
Date: Thu, 10 Nov 1994 16:19:43 +0200 (GMT)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

This is a MIME-encapsulated message

--QAA07917.784477197/santra.hut.fi

The original message was received at Thu, 10 Nov 1994 16:19:57 +0200
from leka.hut.fi [130.233.224.22]

----- The following addresses had delivery problems -----
{listserver-at-anlemc.msd.gov} (unrecoverable error)

----- Transcript of session follows -----
501 {listserver-at-anlemc.msd.gov} ... 550 Host unknown (Name server: anlemc.msd.gov: host not found)

----- Original message follows -----

--QAA07917.784477197/santra.hut.fi
Content-Type: message/rfc822

Return-Path: ert-at-leka.hut.fi
Received: from leka.hut.fi (leka.hut.fi [130.233.224.22]) by santra.hut.fi (8.6.8.1/8.6.7) with ESMTP id QAA07915 for {listserver-at-anlemc.msd.gov} ; Thu, 10 Nov 1994 16:19:57 +0200
Received: (from ert-at-localhost) by leka.hut.fi (8.6.8.1/8.6.7) id QAA110591 for listserver-at-anlemc.msd.gov; Thu, 10 Nov 1994 16:19:44 +0200



dear Nestor,


Microscopists, I will like to participe in the network of the microscopists.

My email: eero.ristolainen-at-hut.fi



Address: Helsinki University of Technology
Center for Chemical Analysis
FIN-02150 Espoo, Finland


Phone: +int 3580 451 2608
FAX: +int 3580 462 373



Dr. Eero O. Ristolainen
Director

--QAA07917.784477197/santra.hut.fi--






From: POSHEL-at-wpo.it.luc.edu
Date: Fri, 11 Nov 1994 08:49:17 -0600
Subject: embedding media & thick sections ..... -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Michael,
Have you considered:
1) freezing sectioning on a sliding (=sledge) microtome
or 2) a vibratome (I've had my best luck with 10% gelatin fixed 16-20 hr
in 10% formaldehyde)
Phil Oshel
poshel-at-luc.edu





From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Fri, 11 Nov 1994 15:59:50 -0600 (CST)
Subject: MSA staff support

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear fellow microscopists,

With budgetary restictions on the sciences these days, many of us are no
longer able to attend the National Microscopy Society of America (MSA)
meeting. We are limited in the ability to present our work, share our
ideas and research, or meet with our colleagues due purely to financial
limitations.

Shams Ghoniem of the Iowa Microscopy Society has suggested that MSA
consider awarding grants to professional and scientific staff similar
to the program currently in place for students. This recommendation,
put forward by Peter Ingram, LAS/MSA president, was met with a lukewarm
response by the MSA council. I believe that MSA would consider such a
proposal if enough support by the MSA membership is provided. MSA
would benefit by increased attendance at meetings, more incentive to
join MSA, as well as in supporting education and professional development
of their members.

I am requesting therfore that letters be sent fo Peter Ingram in support
of this proposal. We need letters from across the country including
letters from faculty who are willing to match an MSA award. These letters
need to be sent by the first of January to be influential at the next MSA
committee meeting.

His address is: Peter Ingram
LAS Director
Research Triangle Institute
Box 12194
3040 Cornwallis Road
Research Triangle Park, NC 27709

E-mail: ingram-at-rcc.rti.org

I thank you for your attentive ear and for your letters!

Kathy Walters
secretary, Iowa Microscopy Society






From: Daniel L. Callahan :      dlc-at-owlnet.rice.edu
Date: Fri, 11 Nov 1994 15:51:53 -0600 (CST)
Subject: AFM tips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I understand that there are sources for particularly sharp AFM tips.
The normal AFM tip has a radius on the order of 10's of nanometers and
isn't very useful for "deep" contours. Does onyone out there know of a
source for sharper tips.

Daniel L. Callahan
Department of Mechanical Engg. and Materials Science
Rice University
dlc-at-owlnet.rice.edu




From: MicroToday-at-aol.com
Date: Sun, 13 Nov 1994 16:22:56 -0500
Subject: Equipment Listing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Group -
Preface: I find myself "semi-retired" (meaning that I have more free
time than most of you), yet interested in microscopy (our newsletter
"Microscopy Today") and rather alert when it comes to data bases.
Result: I am considering building the equipment listing data base that
has been discussed.
My THOUGHT: To build a data base with two sections:
Section 1: A simple listing of members - names, companies, addresses,
tel. & eMail address, etc. To include all interested, not necessarily only
those identified with equipment. Hard copy of this section could, of course,
be easily provided.
Section 2: A listing of equipment categories. I.E., I would expect
SEM, ESEM, etc. to be categories. Each category to be a "field" in which we c
ould list individual products - which can be searched on.
Then, perhaps, to provide the data base on disks for those interested.
Perhaps there is a way that one can access the data base by eMail?
Before I consider getting serious, I have two questions:
1) Would the above solve the need as discussed? Or is there other and/or dif
ferent detail that should be included? Applications?
2) I happen to use "Q&A" as my data base software. In Q&A, items can be
separated in a keyword field with semicolons and then searched on. I can
export to any other data base software. I do not know if other data base
software will allow this keyword search?
Rather than swamping the listserver with comments from the whole
membership, may I suggest that only those who have previously commented on
this subject publish their response to the full readership. Others might
just send their opinions to me.
Assuming that the concept is OK, or close, the next step is to ask your
help in defining categories. I will ask for that assistance should the
project be a "go".
Regards,
Don Grimes, Microscopy Today




From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Mon, 14 Nov 1994 10:33:26 +0000 (GMT)
Subject: Renaming my computer accoun

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Nestor:

My computer address at Cambridge has been changed. It is noe
pe13-at-cus.cam.ac.uk
Will you please make the necessary alteration at your end so I may
continue to get all the good stuff on the MSA file server.

Many thanks

Patrick Echlin
Director, Multi-Imaging Centre






From: Eliesh O'Neil :      eliesh.oneil-at-gtri.gatech.edu
Date: Mon, 14 Nov 1994 09:30:36 -0500 (EST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

unsubscribe eliesh.oneil-at-gtri.gatech.edu
Elizabeth G. (Eliesh) O'Neil
Research Scientist I
Georgia Tech Research Institute
Electro-Optics, Environment, and Materials Laboratory
Materials Analysis Center
925 Dalney Street, Baker 271
Atlanta, Georgia 30332-0827
Phone: (404) 853-0590
FAX: (404) 894-5073
Email: eliesh.oneil-at-gtri.gatech.edu





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Mon, 14 Nov 1994 10:50:55 -0600 (CST)
Subject: Equipment Lists...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


]
Just a reminder....

Please do not forget the User Facilities and equipment
lists and DATABASE that Sandy Silvers has compile and
distributes via her local society. It is called
the MSA TECH FORUM FACILITY DIRECTORY. The last update
that I received from Sandy was dated 7/94. This is
a working DATABASE with the following information...

Facilities ]
Phone Lists
State Listings
Equipment Listings.

The contact phone number listing is : 706-546-3471




From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Mon, 14 Nov 1994 11:49:49 -0500 (EST)
Subject: thanks for shutter info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have called Vincent and Associates at (800) 828-6972 regarding their
products. We are waiting for specifications and pricing via snail mail.
Has anybody experience with operating one of their shutters (e.g. D122 or
iD123) from a serial port, specifically from within NIH-Image? They make
it sound very simple, but we thought we'd find out whether anybody has
experience.

The Ludl system, at first glance, seems far more expensive. Is it
significantly better? And the same question as above regarding the
Vincent shutter.

We're waiting to hear from Scientific Imaging Syatems too.

Thanks again-

Michael








From: MicroToday-at-aol.com
Date: Mon, 14 Nov 1994 12:05:27 -0500
Subject: Equipment Data Base

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Group -
In response to my recent "offer" to develop an appropriate data base, I
note the comment from Nestor where he recommends the data base as developed
by Sandy Silvers. I accept that as a "non-recommendation" to work with me.
I should point out that the "meat" of Sandy's data base is two listings:
1) An "equipment listing" including type/model, year installed, whether
under a service contract, and user charge rates.
2) An "ancillary equipment listing" including quantity, manufacturer, type,
model #s, etc.
There is no way (I BELIEVE) to do a search on specific instruments or
applications - and come up with a list of others with a specific interest.
I should also point out that Sandy's system has sections for:
1) Type of Facility (biology, material science, core univ., etc.
2) List of personnel and users.
3) Other services provided
4) Number of users per year.
Kindly understand that I have NO INTEREST in competing against Nestor and
his wishes - we (including I) owe him too much. And, considering the amount
of work involved, I have no interest in accomplishing a proper, searchable,
data base without the full support of the "management" of the listserver. As
a result, I withdraw my offer - with thanks to you all that have expressed
interest and support.
Don Grimes, Microscopy Today




From: fornof-at-dns.mrs.org (George Fornof)
Date: Mon, 14 Nov 1994 12:22:35 +0500
Subject: unsubscribe Fornof@mrs.org

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe fornof-at-mrs.org

phone 412-367-4003 ext. 650

fax 412-367-4373





From: Alan Pooley :      pooley-at-ahab.rutgers.edu
Date: Mon, 14 Nov 1994 12:54:41 -0500
Subject: Re: Magnetic shielding principles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I know of one case where sheilding was successful. The Field emmission
sem at cell biology, Yale Medical school had all the usual problems until
a layer of (I thought it was brass sheet but am not sure now) metal was applied
on all 6 surfaces of the room at ???? cost.
I don't have his email but Phillip Male (pronounced Mal) at cell biology
yale university is the one to contact, the shielding DID work!
alan pooley marine sci rutgers univ nj




From: Daniel Henne :      henne-at-sfu.ca
Date: Mon, 14 Nov 1994 10:39:10 -0800 (PST)
Subject: EDX Standards

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Via: bham.ac.uk; Mon, 14 Nov 1994 17:15:55 +0000

Does anyone have names and phone numbers for suppliers of standards
for my EDX analysis. I need to know the compositions to see how well
my setup is doing it's work.

Cheers from Sunny Vancouver.
Dan Henne
Simon Fraser University
henne-at-sfu.ca





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Mon, 14 Nov 1994 14:49:01 -0600 (CST)
Subject: Equipment Lists

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Don Grimes & Fellow Microscopy List Readers...

Sorry If I gave the wrong impression. I have no problems
with either Don or Sandy running a DATABASE on equipment.
I just wanted to make sure that Sandy's work over the
last few years was remembered.

Don, your more than welcome to continue to use this forum
for suggestions and/or comments. I apologize if I gave the
wrong impression....


Nestor




From: Michael Boucher :      BOUCHER-at-tcrca.usbm.gov
Date: Mon, 14 Nov 1994 14:04:41 CST
Subject: Old SEM Donation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



As I Mentioned on the net earlier, we have an old JEOL model JSM-2
SEM with a Nuclear Diodes/ EDAX model 508 EDS system, built in 1968.
It has a manual vacuum valving system and the power supply is vacuum
tube based. It was refurbished in the mid 80's and the EDS was
repaired. Although in good operating condition a year ago when last
used, we have no further use for it. We have permission to donate it
to a non-profit charitable institution or University. Anyone wanting
to acquire it would have to come here to Minneapolis, pack it up and
ship it at their own expense. JEOL will not be able to supply parts
for it, so you're on your own there. Only qualified organizations
may apply; no individuals or for-profit businesses. Any requests can
be sent to me at the e-mail address below. If you're seriously
interested, please contact me ASAP. We must move it by the end of
the year.
By the way, thanks for all the advice given to me earlier.

Mike Boucher Boucher-at-tcrca.usbm.gov
*****************************************************************
Michael L. Boucher Sr. Boucher-at-TCRCA.USBM.GOV
Geology-Mineralogy/Chemistry Labs Ph 612-725-4614
Twin Cities Research Center Fax 612-725-4527
U.S. Bureau of Mines Center 725-4500
Department of Interior
5629 Minnehaha Avenue South
Minneapolis, MN 55417-3099
U.S.A.
*****************************************************************





From: howelld-at-egr.msu.edu
Date: Mon, 14 Nov 1994 22:20:37 -0500 (EST)
Subject: Help w/ Positive-Image Transparencies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I need some expert advice on the preparation of positive-image
transparencies. I typically use 35 mm slides in my own
presentations, but one of my collaborators insists on
overhead transparencies. I have seen composite overheads
which appear to have positive images on film attached to them.
So my questions are:

Are these made by exposing light-sensitive film on an
enlarger?

What type of film is used? (Manufacturer, Stock No., Sizes
available)

What are the common exposure times and f-stops on the enlarger?

What are the developer/fixer chemical ratios and developing/fixing
times?

A prompt response would be appreciated since this project was
thrown into my lap on short notice with an even shorter
deadline. BTW, I tried a laser xerox of a print (HRTEM image) onto a
transparency and the results were unacceptable.

Cheers,

David A. Howell
MSM Dept.
Michigan State University
E. Lansing, MI
howelld-at-egr.msu.edu




From: ldm-at-apollo.numis.nwu.edu (L. D. Marks)
Date: Tue, 15 Nov 1994 11:42:16 -0600
Subject: Re: Help w/ Positive-Image Transparencies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411151709.AA29885-at-titian.jeol.com}

Color laser copiers working in B/W mode give very good quality
reproductions, and are far more convenient than printing methods.
Most Kinko's have a copier -- you may have to work with them to make
sure that they give good quality results, and many Universities also
have an appropriate copier.




From: MELSEN :      MELSEN-at-microbio.emory.edu
Date: 15 Nov 1994 16:09:08 EST
Subject: polyester wax supplier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am in search of a US supplier for polyester wax. This product is
produced by a company in the UK.

Please respond directly to my e-mail address

TIA

Skip
melsen-at-MICROBIO.emory.edu




From: kayton-at-ohsu.edu (Robert Kayton,MAC,CROET)
Date: Mon Nov 14 17:10:58 PST 1994
Subject: Re: curling of cryo samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


------------ Original Message -------------



Very often when I am cryo-sectioning , the sections curl so badly
that they are useless. This was the reason for purchasing the static-line
ionizer. This really did not help the curling problem, although it served
other purposes. My sections are mostly polymer-elastomer blends, primarily
nylon based. If you have any suggestions I would be eternally greatful. Also
if anyone has suggestions on how to pick up the samples off of the cryo
knife that would also help. Right now we are using a very fine sewing
needle, but something even thinner yet ridgid would be better


Thank you

Andrea Monisera.
--------- End of Original Message ---------

I have seen the eyelash hairs of Dalmatians (dogs) used for this. they are
forked at the end and work very well for picking up cryosections


Bob Kayton
OHSU
Portland , Oregon

E-mail address kayton-at-ohsu.edu




From: nina allen :      allen-at-wfu.edu
Date: Tue, 15 Nov 1994 16:44:41 -0500 (EST)
Subject: Re: Help w/ Positive-Image Transparencies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Mon, 14 Nov 1994 howelld-at-egr.msu.edu wrote:

You can take your slide to a digitizing scanning device and then feed
that image to print it or to have it printed on an overhead
transparency. This is done in some photo departments.

Nina Allen.

} } I need some expert
advice on the preparation of
positive-image } transparencies. I typically use 35 mm slides in my own
} presentations, but one of my collaborators insists on
} overhead transparencies. I have seen composite overheads
} which appear to have positive images on film attached to them.
} So my questions are:
}
} Are these made by exposing light-sensitive film on an
} enlarger?
}
} What type of film is used? (Manufacturer, Stock No., Sizes
} available)
}
} What are the common exposure times and f-stops on the enlarger?
}
} What are the developer/fixer chemical ratios and developing/fixing
} times?
}
} A prompt response would be appreciated since this project was
} thrown into my lap on short notice with an even shorter
} deadline. BTW, I tried a laser xerox of a print (HRTEM image) onto a
} transparency and the results were unacceptable.
}
} Cheers,
}
} David A. Howell
} MSM Dept.
} Michigan State University
} E. Lansing, MI
} howelld-at-egr.msu.edu
}




From: Glenn Holm :      KARUZIS-at-wccf.mit.edu
Date: Tue, 15 Nov 1994 18:43:46 -0500 (EST)
Subject: Re: polyester wax supplier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Tue, 15 Nov 1994, MELSEN wrote:

} I am in search of a US supplier for polyester wax. This product is
} produced by a company in the UK.
}
} Please respond directly to my e-mail address
}
} TIA
}
} Skip
} melsen-at-MICROBIO.emory.edu
}
I expect there might be enough interest for a general reply to
the list, if anyone knows, since there is an article in the new
Biotechnoques [17(5):846-848] by L.L Richardson and M. Dym about
polyester wax embedding for LM level immuno. They cut at 5-10
micorns at 4 degrees in a cryostat and ethanol dehydrate and air
dry to get the sections to adhere to slides. The supplier they
list is BDH, Poole, UK.
I've never cut the stuff, but their pictures look decent. I have
heard of cutting polyester wax on a sliding microtome with a funnel
of dry ice above the block so the cool vapor keeps the sections
from melting. If anyone has experience with cutting this wax or
info about U.S. suppliers, I'd be interested in hearing about it.

------------------------------------------------------------------
|Glenn Holm *mime mail ok* Internet:karuzis-at-wccf.mit.edu |
|M.I.T Dept. of Brain + Cog. Sci. This VAX doesn't do NeXTmail |
|Cambridge, MA 02139 "Real Neuroscientists don't do gels!" |
------------------------------------------------------------------





From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Fri, 11 Nov 1994 15:59:50 -0600 (CST)
Subject: MSA staff support

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reply... RE} MSA staff support
Reply
from: John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX: (313)936-3352
jfmjfm-at-umich.edu or John_Mansfield-at-mse.engin.umich.edu
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html

This is a reply to the original message posted to the Microscopy List Server
that is included below. These opinions are mine own and do not necessarily
relect those of my employer.

As a University employee who has to watch every dollar, I frequently find it
is expensive to go to conferences. However, I do not think I should get
handouts from MSA. I think that MSA money is better spent encouraging
students (grade school, high school and both undergraduate and graduate
university students) to get interested in various aspects of microscopy.
The outreach programs in schools, the MSA scholoarships and bursaries for
attending our conferences are much more important and have an impact on a
greater number of people than sending a few senior people to the meetings.
There are too many people wanting to milk money out of MSA for their own
benefit, rather than the good of the community as a whole.
As I say, just my "humble" opinion.
John Mansfield.

--------------------------------------

Dear fellow microscopists,

With budgetary restictions on the sciences these days, many of us are no
longer able to attend the National Microscopy Society of America (MSA)
meeting. We are limited in the ability to present our work, share our
ideas and research, or meet with our colleagues due purely to financial
limitations.

Shams Ghoniem of the Iowa Microscopy Society has suggested that MSA
consider awarding grants to professional and scientific staff similar
to the program currently in place for students. This recommendation,
put forward by Peter Ingram, LAS/MSA president, was met with a lukewarm
response by the MSA council. I believe that MSA would consider such a
proposal if enough support by the MSA membership is provided. MSA
would benefit by increased attendance at meetings, more incentive to
join MSA, as well as in supporting education and professional development
of their members.

I am requesting therfore that letters be sent fo Peter Ingram in support
of this proposal. We need letters from across the country including
letters from faculty who are willing to match an MSA award. These letters
need to be sent by the first of January to be influential at the next MSA
committee meeting.

His address is: Peter Ingram
LAS Director
Research Triangle Institute
Box 12194
3040 Cornwallis Road
Research Triangle Park, NC 27709

E-mail: ingram-at-rcc.rti.org

I thank you for your attentive ear and for your letters!

Kathy Walters
secretary, Iowa Microscopy Society



------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;11 Nov 1994 18:33:41 -0400
Received: by judgmentday.rs.itd.umich.edu (8.6.9/2.2)
with X.500 id RAA28989; Fri, 11 Nov 1994 17:31:38 -0500
Received: from AAEM.AMC.ANL.GOV by judgmentday.rs.itd.umich.edu (8.6.9/2.2)
with SMTP id RAA28982; Fri, 11 Nov 1994 17:31:37 -0500







From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Tue, 15 Nov 1994 22:27:36 -0600 (CST)
Subject: Transparencies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1427222913.15887-at-mse.engin.umich.edu}

Reply... RE} } Linking pc to Mac
Reply
from: John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX: (313)936-3352
jfmjfm-at-umich.edu or John_Mansfield-at-mse.engin.umich.edu
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html

I think the easiest way of linking a PC to a Mac over Ethernet would be to
put ethernet cards in both machines (I think you have a Quadra with built in
E-net and so you are all set there. Get a thin net connector for the Quadra
and configure the board for the PC for thin net. If you have an EISA bus PC
I would suggest getting a NE3200 (I think it is) 32 bit card that should give
you the fastest throughput. However if you are on a really tight budget then
you can get a 16bit card from 3Com for less than $200. You can just connect
the two computers with one strip of thin net cable and terminate both ends.
Then I would use public domain or shareware TCP/IP software for each machine.
The Mac OS System 7.5 comes with MacTCP and you can use Fetch, NCSA Telnet
or XferIt (or the commercial program Versaterm) to run ftp on the Mac and
NCSA telnet to run ftp on the PC (there may be other better implementations
of the TCP/IP stack for the PC that are shareware, I am not sure.)
Fetch is really nice, you could put the PC into server mode (run ftpd) and
use the nice interface of Fetch to put or get your files. If you are using
only those two machines you can use pretty much any IP address you like for
the two machines.
If you put them on the Internet though, you will need to get IP addresses
allocated by your local network administrator.
Hope this helps.
John Mansfield
--------------------------------------

} cytogen-at-tigger.jvnc.net wrote:
}
} } Hello-
} } I have a pc-based system to read in digital medical image data. I transfer
} } these as TIFF or Interfiles to a Quadra 950 for processing and display and
} } eventually for submission to the FDA. At the moment I use MacLinkPC to
} } transfer the files over a serial line. At the rate of 57,600 baud, 100Mb
} } of image data takes all night to transfer. What is the EASIEST way to take
} } advantage of the Quadra's built-in Ethernet capability to speed up this
} } transfer? I would prefer a solution which did not include taking a lengthy
} } network administrator's course. Any suggestions will be appreciated.
} }
} } John G Wolodzko
} } (jwolodzko-at-cytogen.com)
}
} Cheapest and fastest might be to use a hard-disk cartridge drive on both
} the pc and Mac, then use sneaker-net!
}
} Roger Pyle

I've used the Syquest cartridge/sneaker-net route and it may be the
fastest. A less expensive solution may be the ethernet and "PhoneNet PC"
software (sold as part of Timbuktu from Farallon). I have an Asante
EN2000+(thin-net) ethernet card on a PC running PhoneNet and a
"friendly-talk" thin-net adapter on the Quadra. I create a "server" share
folder using system 7 Appleshare on the Quadra and use PhoneNet to make the
PC a client on the Quadra's share folder.(You can't make the PC a server
using PhoneNet). With the proper priveliges set up on the Quadra you copy
your files into it's share folder. For easiest running I would call
Farallon and ask them what their default ethernet card is and buy that with
the Timbuktu software. Installation and setup is then straightforward
without "taking a lengthy network administrator's course".
Lots of Luck,
Fran Tanzella
Sr. Chemist, SRI International

------------------ RFC822 Header Follows ------------------
Received: by mse.engin.umich.edu with SMTP;14 Nov 1994 15:13:32 -0400
Received: by truelies.rs.itd.umich.edu (8.6.9/2.2)
with X.500 id OAA05672; Mon, 14 Nov 1994 14:11:17 -0500
Received: from caipfs.rutgers.edu by truelies.rs.itd.umich.edu (8.6.9/2.2)
with SMTP id OAA05661; Mon, 14 Nov 1994 14:11:13 -0500
Received: from [128.18.30.66] by caipfs.rutgers.edu (5.59/SMI4.0/RU1.5/3.08)
id AA02086; Mon, 14 Nov 94 12:30:40 EST
Received: from Tanzella ([128.18.70.30]) by unix.sri.com (4.1/SMI-4.0)
id AA24131; Mon, 14 Nov 94 09:29:39 PST
Message-Id: {Tanzella.1135221855A-at-128.18.30.66}


I haven't used the negative reversal procedure in a few years
to make sheet film into transparencies. It wasn't cost effective
to keep it up and never gave me the quality that I liked.

The method I now use is to use a dye-sublimination printer
to write directly to "transparency" film. This produces photographic
quality output. Of course this means you need to find someone
with a dye-sub printer. Typical cost/print is about $2-$3 for
an 8x10 transparency. The printers themselves run about $10K.

If you have a good laser-printer (at least 300 dpi) and a graphics/image
processing program that does reasonable dithering (Adobe Photoshop for
example) then you can print to "laser printer transparency" film.
This give marginally acceptable prints but I wouldn't want to use these
unless there was no other alternative. The better laser printers
support a "photo-grade" mode (the name varies depending upon the manufacturer)
which greatly improves the gray-scale dithering pattern, but it still won't
compete with dye-subs (in my opinion).

The very high end printers (~2700+dpi)
like the Linotype Imagesetters give results which compare with
dye subs. These are typically found in graphic arts departments.
If you look with an eye piece you can still see the dither pattern
but it's pretty small.

There are also the ink-jet printers which depending upon the graphic
do okay jobs. I tend not to use these for continuous tone images but
prefer them on solid fill graphics. Mainly due to their lower cost/print
~$1 for an 8x10.

But all of this assumes that you have the images/slides already
in some computer and ready for printing. If not then you'll have
to go across to your graphic arts department and have them
shoot an 8x10 negative and do the reversal for you. Last time
I checked the price for this (here at ANL) was about $12/transparency.

This likely doesn't solve your problems but at least you can ask
around and see if the printers are handy. I would be surprized
if there were none on your campus.

--Nestor




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Wed, 16 Nov 1994 11:46:07 -0600 (CST)
Subject: System Crashes & Doubles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sorry Subscribers, I'm still having a system problem.
You will likely continue to get doubling of some
messages that donot clear the queues before a crash
happens. I'm looking into the problem.

Nestor
:-(

-------




From: jan.czernuszka-at-materials.oxford.ac.uk
Date: Wed, 16 Nov 1994 16:43:29 +0000
Subject: System Crashes & Doubles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: jan.czernuszka-at-materials.oxford.ac.uk





From: Bonnie Davis - Kennametal Inc. :      bld_kmt-at-prlc.org
Date: Wed, 16 Nov 1994 12:46:14 -0500 (EST)
Subject: Messages in triplicate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am receiving messages in triplicate!!! Is there a way to correct this?

Thanks




From: Michael Rock :      merock-at-u.washington.edu
Date: Wed, 16 Nov 1994 12:11:46 -0800 (PST)
Subject: Philips 300G TEM (blue light special)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Via: bham.ac.uk; Wed, 16 Nov 1994 13:26:58 +0000

this one's gotta go!

We are looking for a good home for a fully operational Philips 300G TEM.
New HT tank stable operating at 20,40,60,80, and 100 kV.
Both biological stage and goneometer stage.
Several specimen holders (single-tilt, double-tilt, tilt-rotate).
Planning on surplusing soon (Univ. of Wash., Seattle, WA).

Interested parties contact by email: merock-at-u.washington.edu

-Mike Rock
EM Lab Manager





From: norenbur-at-onyx.si.edu (Jon L. Norenburg)
Date: Wed, 16 Nov 1994 10:09:48 -0500
Subject: Re: polyester wax supplier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English

} I am in search of a US supplier for polyester wax. This product is
} produced by a company in the UK.
}
} Please respond directly to my e-mail address
} melsen-at-MICROBIO.emory.edu

Try Polysciences, Warrington, PA or Aldrich Chemical Company, Milwaukee, WI.
I haven't used polyester in years, but both stocked it as of 1988. Aldrich
sold 5 gal and up quantities and a wide range of mol wts; I had gotten two
1-gal samples from them that lasted me years. (See also my paper - 1987,
JEM Technique 6:35-41).
--Jon


Jon L. Norenburg {norenbur-at-onyx.si.edu}
Invertebrate Zoology, National Museum of Natural History
Smithsonian Institution, Washington, DC 20560
Voice 301-238-3508, Fax 301-238-3361






From: Carl Henderson :      chender-at-umich.edu
Date: Wed, 16 Nov 1994 11:47:28 -0500 (EST)
Subject: SEM: Rebuilding Hitachi gun

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Return-Path: {dieguez-at-iris1.fae.ub.es}
Received: from oxmail.ox.ac.uk by vax.ox.ac.uk (MX V4.1 VAX) with SMTP; Wed, 16
Nov 1994 00:39:56 +0000
Received: from aaem.amc.anl.gov by oxmail.ox.ac.uk with SMTP (PP) id
{12213-0-at-oxmail.ox.ac.uk} ; Wed, 16 Nov 1994 00:39:34 +0000

Has anyone had experience rebuilding an SEM electron gun?

Our Hitachi S-570 SEM has had emission current fluctuations which I
traced to the gun cable-gun connection inside the gun. Trouble is,
the connections are surrounded by a white, opaque (silicone?) potting
compound. When I press down on the potting compound, my emission current
troubles disappear, thus I know that the connection is shaky inside.

I'm guessing that I can carefully cut away the old potting compound,
expose and repair the cable to gun connections, and finally repot the
gun. Not knowing exactly what I am getting into, I thought I would ask
for some net wisdom before the first slice.

BTW, the instrument is 9 years old and a new gun and cable assembly runs
about $5K from Hitachi. There's my incentive!

Thanks,

Carl

Carl Henderson
Electron Microbeam Analysis Laboratory
University of Michigan
2005 C.C. Little Bldg.
Ann Arbor, MI 48109-1063 USA
-----------------------------------------------------
(313) 936-1550 (voice) **** Next time, ****
(313) 763-4690 (FAX) *** take the ***
chender-at-umich.edu (e-mail) **** train! ****
----} (While you still can!)
-----------------------------------------------------




From: Carl Henderson :      chender-at-umich.edu
Date: Wed, 16 Nov 1994 17:05:06 -0500 (EST)
Subject: Re: EDX Standards

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



On Mon, 14 Nov 1994, Daniel Henne wrote:

} Does anyone have names and phone numbers for suppliers of standards
} for my EDX analysis. I need to know the compositions to see how well
} my setup is doing it's work.
}

Another source is:
Cannon Microprobe
1041 NE 100th St.
Seattle, WA 98125
(206) 522-9233
Attn: Bart Cannon

They can make a SEM mount with your choice of standards from a list of
hundreds of minerals/materials. The prices are also reasonable.

Disclaimer: I have not purchased the standards, nor do I have any
association with Cannon Microprobe. Just passing along some info I
recently came across.

Carl Henderson
Electron Microbeam Analysis Laboratory
University of Michigan
2005 C.C. Little Bldg.
Ann Arbor, MI 48109-1063 USA
-----------------------------------------------------
(313) 936-1550 (voice) **** Next time, ****
(313) 763-4690 (FAX) *** take the ***
chender-at-umich.edu (e-mail) **** train! ****
-----------------------------------------------------





From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Wed, 16 Nov 1994 07:19:51 -0600 (CST)
Subject: MSA support

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ron,

Thank you for putting forward your comments, so that I may effectively
clairify my position.

I strongly feel that University Professional and Scientific staff (P&S)
are often overlooked for their contributions to microscopy. Being a part
of the University community, I can only speak from that perspective. I do
not know how easy or difficult it is for staff from industry or goverment
receive support from their employers. I do know that University P & S
staff as a rule have very little control over use of grant money for
travel, and also have much smaller salaries with which to fund themselves
to go to meetings. I do not suggest that MSA pick up the tab for all
expenses, but funding from a $10,000 source could go a long way towards
improving someone's chances of obtaining matching funds from their
employers or other university sources. As an example, there could be
twenty-five $400 awards. I may be accused of thinking small, but what I
originally had in mind was $2000 total for the program (perhaps five $400
awards) with a simple waiving of registration fees for another 5 people.
Why couldn't we try this as apiolet program for one year? If no one
applies it hasn't cost MSA a cent.

As Microscopists we cannot afford to be stagnant. There has to be
continuing education so that we are able to provide meaningful assistance
to our employers, and the many opportunities available at the annual MSA
meeting are the best source for this nessecary growth in ourselves.

Repectfully yours,

Kathy Walters






From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Wed, 16 Nov 1994 08:43:31 +0000 (GMT)
Subject: Re: curling of cryo samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk


Return-Path: {dieguez-at-iris1.fae.ub.es}
Received: from oxmail.ox.ac.uk by vax.ox.ac.uk (MX V4.1 VAX) with SMTP; Wed, 16
Nov 1994 02:35:10 +0000
Received: from aaem.amc.anl.gov by oxmail.ox.ac.uk with SMTP (PP) id
{16027-0-at-oxmail.ox.ac.uk} ; Wed, 16 Nov 1994 02:34:50 +0000
!temhelp {microscopy-at-aaem.amc.anl.gov}

When cutting cryo-sections it is necessary to have an anti-roll plate on
the knife. The best probe to use to pick up frozen sections, is an eye
lash stuck to an applicator stick. An anti static gun may help to reduce
the charge build up in the block-section-knife region.
If all else fails read the appropriate chapter in "Low Temperature
Microscopy and Analysis" by Patrick Echlin, Published by Plenum in 1992.

Good luck with one of the most frustrating activities in cryomicroscopy

Patrick Echlin
Director, Multi-Imaging Laboratory
School opf Biological Sciences
CambridgeOn
Mon, 14 Nov -1 kayton-at-ohsu.edu wrote:

} ------------ Original Message -------------
}
}
}
} Very often when I am cryo-sectioning , the sections curl so badly
} that they are useless. This was the reason for purchasing the static-line
} ionizer. This really did not help the curling problem, although it served
} other purposes. My sections are mostly polymer-elastomer blends, primarily
} nylon based. If you have any suggestions I would be eternally greatful. Also
} if anyone has suggestions on how to pick up the samples off of the cryo
} knife that would also help. Right now we are using a very fine sewing
} needle, but something even thinner yet ridgid would be better
}
}
} Thank you
}
} Andrea Monisera.
} --------- End of Original Message ---------
}
} I have seen the eyelash hairs of Dalmatians (dogs) used for this. they are
} forked at the end and work very well for picking up cryosections
}
}
} Bob Kayton
} OHSU
} Portland , Oregon
}
} E-mail address kayton-at-ohsu.edu
}




From: kelley-at-mrc.com
Date: Thu, 17 Nov 1994 09:53:23 -0500 (EST)
Subject: Re: Transparencies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411161648.AA13779-at-riker.ml.wpafb.af.mil}

Nestor writes:
*
* The method I now use is to use a dye-sublimination printer
* to write directly to "transparency" film. This produces photographic
* quality output. Of course this means you need to find someone
* with a dye-sub printer. Typical cost/print is about $2-$3 for
* an 8x10 transparency. The printers themselves run about $10K.
*
There is a dye-sub printer (PrimeraPro) available from:
Fargo Electronics
7901 Flying Cloud Dr.
Eden Prairie, MN 55344
(800) 258-2974
which lists for $1895.00 with 8X10 prints costing about $3-$4
each. There is also a 4X5 format available at a little over $1
per print, you just have to switch the ribbon and feed tray.
I have tested this printer for wax thermal transfer,
dye-sub prints and dye-sub transparencies. Although it may not
be the same quality you will find form the $10K printer, it appears
to me to give you } 95% of the performance for less than 20% of the
price.

Keith A. Kelley
Miles Research Center
kelley-at-mrc.com





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Thu, 17 Nov 1994 10:05:37 -0600 (CST)
Subject: Isaac???

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



There is a software package written by David Bright called
MacLispX which might be what your thinking about. I've never
heard of Isaac. MacLispX is available downloadable from
the EMMPDL FTP site (WWW.AMC.ANL.GOV) check out the
ANL Shareware/Mac Shareware/Imaging/MacLispX area

Nestor J. Zaluzec




From: John Phelps :      phelps-at-ENH.NIST.GOV
Date: Thu, 17 Nov 1994 08:21:03 -0700 (MST)
Subject: Re: ISAAC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For information about the ISSAC system, contact Michael Postek at NIST,
Gaithersburg. 301-975-2299

John


}
}
} There is, I think, a software package from NIST called ISAAC.
} It is supposed to handle images with more pixels and more bits
} than other programs for acquiring SEM images.
} 1 Does anyone know where to get more information on this?
} 2 Does anyone have experience using it?
}
} If you want a giggle, read the piece in New Scientist (November 5 1994,
} page 21), which is where I heard about it.
}
} Alwyn Eades
} Alwyn Eades
} Center for Microanalysis of Materials
} University of Illinois


John Phelps
NIST - Boulder, CO
303-497-7570




From: rms-at-vax.ox.ac.uk
Date: Thu, 17 Nov 1994 16:54:24 +0000
Subject: November 1994 issue of the Journal of Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk
JTERLET-at-CEMMA.ADELAIDE.EDU.AU, microtoday-at-aol.com, ROSS.MACKENZIE-at-OX.AC.UK,
MICROSCOPY-at-AAEM.AMC.ANL.GOV
Message-ID: {009879C4.579956F3.15-at-vax.ox.ac.uk}

ABSTRACTS FOR THE JOURNAL OF MICROSCOPY, VOLUME 176 PART 2, NOVEMBER 1994.
**************************************************************************

Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 85-98.

High performance X-ray detection in a new analytical electron
microscope

C E Lyman, J I Goldstein, D B Williams, D W Ackland, S von
Harrach, A W Nicholls & P J Statham, Materials Science &
Engineering, Lehigh University, Whitaker Laboratory, 5E Packer
Avenue, Bethlehem PA 18015-3195, USA


SUMMARY

X-ray detection by energy-dispersive spectrometry in the
analytical electron microscope (AEM) is often limited by low
collected X-ray intensity (P), modest peak-to-background (P/B)
ratios, and limitations on total counting time (t) due to
specimen drift and contamination. A new AEM has been designed
with maximization of P, P/B and t as the primary considerations.
Maximization of P has been accomplished by employing a field-
emission electron gun. X-ray detectors with high collection
angles, high-speed beam blanking to allow only one photon into
the detector at a time, and simultaneous collection from two
detectors. P/B has been maximized by reducing extraneous
background signals generated at the specimen holder, the
polepieces and the detector collimator. The maximum practical t
has been increased by reducing specimen contamination and
employing electronic drift correction. Performance improvements
have been measured using the NIST standard Cr thin film. The 0.3
steradian solid angle of X-ray collection is the highest value
available. The beam blanking scheme for X-ray detection provides
3-4 times greater throughput of X-rays at high count rates into
a recorded spectrum than normal systems employing pulse-pileup
rejection circuits. Simultaneous X-ray collection from two
detectors allows the highest X-ray intensity yet recorded to be
collected from the NIST Cr thin film. the measured P/B of 6300
is the highest level recorded for an AEM. In addition to
collected X-ray intensity (cps/nA) and P/B measured on the
standard Cr film, the product of these can be used as a figure-
of-merit to evaluate instruments. Estimated minimum mass fraction
(MMF) for Cr measured on the standard NIST Cr thin film is also
proposed as a figure-of-merit for comparing X-ray detection in
AEMs. Determinations here of the MMF of Cr detectable show at
least a threefold improvement over previous instruments.


Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 99-
109.

The preparation of cryosections from plant tissue: an alternative
method appropriate for secondary ion mass spectrometry studies
of nutrient tracers and trace metals

Dennis B Lazof, Jack G Goldstein, Thomas W Rufty, Cynthia Suggs
& Richard W Linton, USDA-ARS, Crops Research Laboratory, PO Box
1168, Oxford NC 27565, USA


SUMMARY

A method involving cryostat sectioning (10 micrometre thickness)
and freeze-drying is presented for the preparation of plant
tissue for microanalytical studies. The method is well suited for
semi-quantitative imaging by secondary ion mass spectrometry
(SIMS) and offers significant advantages over bulk freeze-dried
or freeze-substitution preparations. Segments of corn or soybean
root (5mm) are quench-frozen, embedded externally, sectioned in
a cryostat (10 micrometre), pressed onto ultrapure Si and slowly
freeze-dried. Images of these sections with secondary electron
microscopy and SIMS indicated good morphological preservation.
It was possible to section tissues of a wide developmental range,
as well as roots varying sixfold in diameter. SIMS images are
presented which demonstrate the ability to detect and localize
nutrient tracers, such as Rb+, following brief exposures (10 min)
to the intact plant. Likewise, a toxic metal (Al) was localized
in root tissue after brief exposure ( {1 day) of the intact plant
root to micromolar external concentrations. Elemental
redistribution during processing was minimal, as demonstrated
most explicitly by the lack of movement of loosely bound Ca from
the outer cell walls into the adjacent embedding material.
Preservation of compositional differences between cellular
content and cell wall was supported by a semi-quantitative
treatment of SIMS images.


Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 110-
120.

Vitrification of aqueous suspensions from a controlled
environment: an improved plunging device

B J Battersby, J C W Sharp, R I Webb & G T Barnes, Department of
Chemistry, The University of Queensland, Queensland 4072,
Australia

SUMMARY

In the process of vitrifying aqueous suspensions for cryo-
transmission electron microscopy, water is solidified without
crystallization. Vitrification can be achieved by rapidly
plunging an aqueous thin film into a liquid cryogen. The
preparation of aqueous thin films prior to vitrification must be
performed in an environmental cabinet at controlled temperature
and humidity in order to prevent evaporation and temperature-
induced phase changes in the thin film. The device described here
incorporates several important features which make the apparatus
simpler and more convenient to use than similar devices described
in the literature. One of these features includes the use of a
totally enclosed environmental cabinet in which the grid, sample,
micropipette and absorbent paper are equilibrated before thin-
film preparation. Other features include a cryogen dewar on a
swing arm for easy refilling, a guillotine shutter which is used
to trigger the plunger electrically and a semi-automatic system
which facilitates rapid transfer of the vitrified specimen from
liquid propane to liquid nitrogen for storage and reduces
handling of the specimen. To demonstrate the utility of the
device, results showing the influence of temperature on the
morphology of phospholipid vesicles are presented. A commercial
cryotransfer apparatus (which is used for the transportation of
the vitrified specimen to the electron microscope cold stage) has
been modified to reduce the possibility of reversion of the
vitreous phase to the crystalline ice phases.


Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 121-
131.

An atomic force microscope for cytological and histological
investigations

Tullio Mariani, Antonio Musio, Carlo Frediani, Isabella Sbrana
& Cesare Ascoli, Istituto di Biofisica - CNR, Via Lorenzo 26,
56127 Pisa, Italy


SUMMARY

An atomic force microscope (AFM) specifically designed for
cytological and histological studies and able to operate on the
same scale of the highest optical magnification is described. The
AFM is a non-invasive instrument; it operates on samples which
do not require any kind of treatment and it can produce
information that supplements and completes the information given
by traditional microscopical methods. The apparatus has been used
to image fixed human chromosomes and to investigate the action
of trypsin during the staining for banding. First results showed
that banding patterns very similar to G-banding pre-exist to
staining and to trypsin treatment in human metaphase chromosomes,
and that the trypsin treatment induces a structural collapse in
the chromatin. The instrument was also used on thin sections of
plant tissue and gave promising results. Experience confirmed
that AFM is a suitable tool for this kind of investigation, and
proved the importance of developing AFMs specifically designed
for routine use in cytology and histology, conceived for non-
specialized users.



Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 132-
142.

Unstained and in vivo fluorescently stained bacterial nucleoids
and plasmolysis observed by a new specimen preparation method for
high power light microscopy of metabolically active cells

Eduard Kellenberger & Cornelia Kellenberger van der Kamp,
Institut de Genetique et Biologie, Microbienne (IBGM), Universite
de Lausanne, Rue Cesar Roux 19, CH-1005 Lausanne, Switzerland


SUMMARY

Microscope slides were coated with a layer of gelatin, the
thickness of the gelatin increasing linearly along the long axis.
The bacterial suspension is applied to the dried gelatin and
covered by a coverslip. The medium is absorbed by the gelatin and
thus the cells applied against the coverslip. By this method,
cultures of concentrations below 100 000 000 cells/ml provide
statistically relevant numbers for observation without prior
concentration steps. It is easier to apply than the existing
methods for the observation of bacterial nucleoids by phase
contrast imaging. Because the cells are maintained in growing
conditions the method is useful for the vital fluorescence DAPI-
staining of various bacterial species and for observations of
plasmolysis and its reversal at different physiological
conditions and extracellular osmolalities. The previously
generally assumed view that the plasmolytic changes of the cell
morphology are immediate upon the hyperosmotic shock and are
rapidly repaired when the cell is able to metabolize actively was
confirmed; this is in contrast to some recent claims.



Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 143-
151.

Fibre optic scrambling in light microscopy: a computer simulation
and analysis

Frederick B Reitz & Len Pagliaro, Center for Bioengineering
WD-12, University of Washington, Seattle, Washington 98195, USA


SUMMARY

Optical fibres bent in two mutually perpendicular planes have
proven useful for randomizing illumination in light microscopes.
These optical scramblers can increase the resolution and/or
contrast obtained with several modes of light microscopy. Here
computer simulations are used to investigate several factors
affecting light randomization in curved optical fibres in order
to further the theoretical basis for scrambler design. Light
passing through 90 degree bends of optical fibre of varying radii
of curvature was modelled by ray tracing in two dimensions, and
scrambling mechanisms were observed. The effects of varying the
position and angle of entry of light on the phase and direction
of propagation of emergent light were determined. It was found
that (a) thorough scrambling does not necessarily require high
numerical aperture entry of light into the fibre, (b)
considerable order persists after a single 90 degree bend of an
idealized fibre and (c) a higher degree of scrambling (at the
cost of transmission efficiency) is achieved in more tightly
curved fibres. The pathlength variations introduced by scrambling
proved smaller than typical laser coherence lengths, requiring
temporal scrambling (vibrating the fibre).



Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 152-
157.

Edge detection, three-dimensional cell boundary reconstruction
and volume and surface area estimation from differential
interference contrast images

C M Kenyon, M Yanai & P T Macklem, Meakins Christie Laboratories,
McGill University, Montreal, Canada H2X 2P2


SUMMARY

Three-dimensional (3-D) cell morphology is important for the
understanding of cell function and can be quantified in terms of
volume and surface area. Differential interference contrast (DIC,
or Nomarski) imaging can enable cell edges to be clearly
visualized in unstained tissue due to the slight difference in
refractive index between aqueous media and cytoplasm. DIC is
affected in only one direction - the direction of the optical
shear. A 1-D edge detector was used in that direction with a
scale length equal to that of an in-focus edge to highlight cell
boundaries. By comparison with the signal from the edge detector
on an out-of-focus slice, the in-focus slices could be segmented
and, after noise suppression, cell outlines obtained. A voxel
paradigm was used to calculate cell volume and differential
geometry was used for surface area estimation. We applied this
approach to obtain 3-D information by optical sectioning of
motile Amoeba proteus.


Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 158-
166.

Nuclear diffuseness as a measure of texture: definition and
application to the computer-assisted diagnosis of parathyroid
adenoma and carcinoma

A J Einstein, J Barba, P D Unger & J Gil, The Lillian & Henry M
Stratton -, Hans Popper Department of Pathology, Box 1194, One
Gustave L Levy Place, New York NY 10029, USA


SUMMARY

A measure of texture, the nuclear diffuseness, was formulated for
use in biological classification, and specifically to
characterize quantitatively chromatin texture. Nuclear
diffuseness corresponds to the amount of local intensity
variation in the digitized image of a nuclear profile. As a
setting in which to test the efficacy of nuclear diffuseness as
a diagnostic tool, the identification of parathyroid adenoma and
carcinoma was considered. Digitized images of sections of
parathyroid chief cell nuclei were obtained from 16 biopsies, and
the nuclear diffuseness, as well as other morphometric
descriptors, were computed. With just the average nuclear
diffuseness and average nuclear profile area, jackknife (leave-
one-out) classification using an artificial neural network was
able to diagnose correctly and unambiguously the condition
(normal, parathyroid adenoma, or parathyroid carcinoma) in 15 of
16 cases. In one case, the neural network assigned a higher
weight to the correct diagnosis, but was unable to distinguish
between normal and adenoma conclusively.


Journal of Microscopy, Vol. 176, Pt. 2, November 1994, pp. 167-
177.

Estimation of the transport properties of polymer composites by
geodesic propagation

R Bremond, D Jeulin, P Gateau, J Jarrin & G Serpe, Centre de
Geostatistique, Ecole Nationale Superiere des Mines, 35 Rue
Saint-Honore, F 77305 Fontainebleau, France


SUMMARY

The purpose of this study is to estimate the transport
coefficients (diffusion, permeability) of polymer composites from
two-dimensional images of these media (polyethylene/polyamide
alloys) obtained by the scanning electron microscope. The main
characteristics of the media investigated are a wide
morphological variability (nodular and lamellar media) and
transport properties that are very different according to the
phase. The computation tools used must accordingly have a wide
range of validity. The idea of the proposed algorithm is to
calculate the distance from an edge of the image, with a modified
distance according to the phase crossed. The results obtained are
compared with those obtained by finite difference integration of
Fick's laws, and with permeability measurements of hydrocarbons
in the materials.








From: tivol-at-tethys.ph.albany.edu
Date: Thu, 17 Nov 1994 13:04:36 EST
Subject: Re: Information on Electron Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tim Prusnick asked in a post to sci.techniques.microscopy about info on EM.
Dear Tim,
For technical aspects on how EM's work and as a referrence, I reccom-
mend Principles of Electron Optics by P.W. Hawkes and E. Kasper from Academic
Press (St Edmundsbury Press Ltd. in the UK).
Yours,
Bill Tivol




From: HOWEY-at-UWYO.EDU
Date: Thu, 17 Nov 1994 15:14:59 -0700 (MST)
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


UNSUBSCRIBE MICROSCOPY RICHARD HOWEY




From: Damon Heer :      DLH-at-fei2.feico.com
Date: Thu, 17 Nov 1994 17:09:48 -0800
Subject: Aperture vendors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings!

I am in search of alternative vendors for electron optical apertures.

What I am particularly looking for is a 2 mm diameter disc / 15
micron aperture made of platinum.

Thanks in advance!
Damon L. Heer

FEI Company
7451 N.E. Evergreen Parkway
Hillsboro, OR 97124-5830

Phone (503) 640-7582
fax (503) 640-7509
email dlh-at-feico.com




From: tony-at-emu.su.oz.au (Tony Romeo)
Date: Fri, 18 Nov 1994 13:34:03 +0100
Subject: VACATION

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411180227.AA00396-at-ELECTRON.emu.su.OZ.AU}
X-Sender: tony-at-electron.emu.su.oz.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

To:

UNSUBSCRIBE MICROSCOPY

Tony Romeo Internet: tony-at-emu.su.oz.au
Electron Microscope Unit
Madsen Building F09 Telephone: +61 2 351 2351
The University of Sydney Facsimile: +61 2 552 1967
Sydney, NSW 2006
Australia







From: tony-at-emu.su.oz.au (Tony Romeo)
Date: Fri, 18 Nov 1994 13:37:37 +0100
Subject: VACATION

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411180231.AA00410-at-ELECTRON.emu.su.OZ.AU}
X-Sender: tony-at-electron.emu.su.oz.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Usubscribe Microscopy tony-at-emu.su.oz.au (Tony Romeo)

Tony Romeo Internet: tony-at-emu.su.oz.au
Electron Microscope Unit
Madsen Building F09 Telephone: +61 2 351 2351
The University of Sydney Facsimile: +61 2 552 1967
Sydney, NSW 2006
Australia







From: PHMOULDK-at-usthk.ust.hk
Date: 18 Nov 1994 18:00:56 +0800
Subject: Training courses.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greeting's

I would to know if there are any training courses specifically for the
preparation of biological materials for examination in a SEM and TEM.

Sorry, I cannot be specific about the type of samples, I just need to know
about general techniques (embedding, fixing, staining etc).

Are there any held by the RMS for example?

As I am in materials characterisation centre, the staff (including me!) have
little experience of the preparation techniques. In particular hands on
training would be a must.



Keith Moulding

Materials Characterisation And Preparation Centre,
Hong Kong University of Science and Technology,
Clear Water Bay,
Kowloon,
Hong Kong.






From: bloody-at-eucmvx.sim.ucm.es
Date: Fri, 18 Nov 1994 18:26:17 +0200
Subject: unsubscribe, PLEASE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


UNSUBSCRIBE MICROSCOPY




From: bloody-at-eucmvx.sim.ucm.es
Date: Fri, 18 Nov 1994 18:27:33 +0200
Subject: UNSUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


UNSUBSCRIBE MICROSCOPY Bloody-at-eucmvx.sim.ucm.es




From: ZWANG-at-utkvx.utk.edu
Date: Fri, 18 Nov 1994 12:31:16 -0500 (EST)
Subject: Used TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






From: T. Xu :      tzxu-at-cml.berkeley.edu
Date: Fri, 18 Nov 1994 10:57:33 -0800
Subject: change of address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please change my email address to tzxu-at-epafi.ME.Berkeley.edu
Thanks
T.Xu




From: bloody-at-eucmvx.sim.ucm.es
Date: Fri, 18 Nov 1994 18:27:33 +0200
Subject: UNSUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Received: from AAEM.AMC.ANL.GOV [146.139.72.3] by beta.ciit.org
with SMTP-OpenVMS via TCP/IP; Fri, 18 Nov 1994 14:37 +0500

UNSUBSCRIBE MICROSCOPY Bloody-at-eucmvx.sim.ucm.es





From: Duke, Steve :      DUKE-at-uthscsa.edu
Date: Fri, 18 Nov 1994 16:17:03 -0600
Subject: Staining of polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Our interests are to descriminate the interdiffusion of methacrylate =
resins into mineralized substrates using TEM. At present, we are unable =
to distinguish resin interdiffusion depth versus epoxy embedding media =
following decalicification. Ideas are to find specific stains for the =
resins, or, maybe label the methacrylate with a reactive colloidal gold. =
Any suggestion on how to descriminate these phases would be appreciated =
very much. Help needed,,,thank you




From: Fred Schamber :      73751.3677-at-compuserve.com
Date: 18 Nov 94 15:17:40 EST
Subject: EM - what major contributions?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings:

A member of our organization has been asked to make a presentation related to EM
in education. Knowing that this is a topic near to the hearts of many on this
list, we thought we might ask for some input.

(1) As part of the presentation, we intend to list some of the principal areas
where EM has made important contributions. Specifically, we're trying to
identify areas where a discovery of large societal significance owes principally
to EM. We have a few ideas, but are looking for others. Suggestions ?

(2) We're looking for the names of people who might have had experience in using
EM in educational programs (as opposed to research) at the secondary and/or
college levels, or have a strong interest in promoting same.

People with suggestions can contact me directly or via this list-server.

Thanks,

Fred Schamber
Instruments Division
RJ Lee Group







From: Brian Oates :      oates-at-unixg.ubc.ca
Date: Sat, 19 Nov 1994 15:30:24 -0800 (PST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Return-Path: {dieguez-at-iris1.fae.ub.es}
Received: from AAEM.AMC.ANL.GOV by INS.INFONET.NET (MX V4.1 AXP) with SMTP;
Wed, 16 Nov 1994 18:20:11 CST

unsubscribe microscopy




From: Duke, Steve :      DUKE-at-uthscsa.edu
Date: Sun, 20 Nov 1994 05:16:54 -0600
Subject: EM of Polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Our interests are to descriminate the interdiffusion of methacrylate
resins into mineralized substrates using TEM. At present, we are unable =
to distinguish resin interdiffusion depth versus epoxy embedding media =
following decalicification. Ideas are to find specific stains for the =
resins, or, maybe label the methacrylate with a reactive colloidal gold. =
Any suggestion on how to descriminate these phases would be appreciated =
very much. Help needed,,,thank you




From: Dave Calvert :      73363.1014-at-compuserve.com
Date: 20 Nov 94 18:22:21 EST
Subject: Just before I shot out the CRT...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry - couldn't seem to send my real message so composed this a test. did
you see T.B.'s joke - here I'll stick them on here

A joke I read: What do you call the ugly piece of flesh on the end of a
penis?

A man.

A joke I made-up, in case you had to guess:

What do you call the Speaker of the House after a special,
behind-closed-doors meeting with Hillary Clinton?

Neutered Gingrich





From: M.Dickson-at-unsw.edu.au (Mel Dickson)
Date: Mon, 21 Nov 1994 13:51:17
Subject: Re: Gold sputter targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-aaem.amc.anl.gov

In article YBerta-at-ms-mail.chemse.gatech.edu writes:
} }
} We need to replace the gold target on our ISI sputter coater; SPI is out of
} the targets right now and Electron Microscopy Sciences doesn't carry targets
} for ISI.
}

Any competent local gold worker can make you a target much cheaper than the
specialist suppliers. Many of Sydneys jewellers have manufacturing capacities
and can easily make me a target any specified diameter and thickness. Better
still, they will accept an old perforated target and add gold to restore a
solid sheet. Gold palladium however is too hard and has too high a melting
point for local manufacturers to rework.




From: M.Dickson-at-unsw.edu.au (Mel Dickson)
Date: Mon, 21 Nov 1994 14:19:44
Subject: Re: Ernest Fullam Tensile stage on S360

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-AAEM.AMC.ANL.GOV

Hello out there,

We just had a 100 lb tensile stage delivered from Ernest Fullam to fit on our
Leica/Cambridge S360. Is there any one out there who has one fitted on an
S360? We need some feedback on practical fitting. The main problem is the
plate that adapts the drive motor to the left side port has no flat ground on
its edge to clear the door hinge. Our workshop can easily grind the flat, but
I'd like a clue where the (eccentrically located) motor should be located.
Looks as if 12 O'clock to 3 O'clock might suit. Any other feedback on
operation will be welcome as the manual is a minimalist document.




From: timonf-at-earth.ruu.nl (Timon Fliervoet)
Date: Mon, 21 Nov 1994 14:23:41 +0100
Subject: Re: In Situ Experiments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Randi,

An oldi, but very nice article is written by Naziri et al. 1972:
In situ superplasticity experiments in the 1 million volt electron microscope.
J. Microscopy, 97, 229 - 238

Hopes this helps.

Cheers, Timon

} Hello!
}
} I wonder if anyone can help me:
}
} I just got the title of a lecture I need to give in
} connection to my Ph.D. 'defence' in 14 days..:
}
} "In Situ Experiments in the Transmission Electron Microscope"
}
} I have found some references in the ICEM-94 proceedings
} and some late numbers of Ultramicroscopy....
}
} -Does anyone have some other references?
}
} -Are you working with this, I am very interested in getting
} some ideas!!!
}
} -What are the main problems? -What can be studied?
}
} etc...
}
} You may respond directly to me or through the news-group!
}
} I am waiting..
}
} Randi.
}
} email: randih-at-imf.unit.no

------------------------------------------------
Timon Fliervoet, Timonf-at-earth.ruu.nl, Faculty of Earth Sciences, Department
of Geology, Utrecht University, P.O. Box 80.021 3508 TA Utrecht, the
Netherlands.
tel: ++31 30 - 535054, fax: ++31 30 - 537725






From: fellers-at-mailer.martech.fsu.edu (Tom Fellers)
Date: Mon, 21 Nov 1994 10:06:57 -0800
Subject: In situ experiments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Randi,
A recent issue of the Materials Research Society bulletin (MRS Bulletin,June
1994,Volume XIX, No.6) is devoted almost entirely to materials science
research now being done using in situ electron microscopy.

________________________________________________________________
Tom Fellers
Florida State University
Center for Materials Research and Technology
Tallahassee, FL 32306-4000
(904) 644-6559 fellers-at-phy.fsu.edu





From: Dave DeFily :      defily-at-tam2000.tamu.edu
Date: Mon, 21 Nov 1994 10:03:07 -0600
Subject: Mac video "field" grabber

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a video board for the Mac which will allow acquisition of
the next video field after an external trigger? The boards I've seen (Scion
LG-3 and Neotech) will allow display of a single field, but even or odd has to
be specified. What I need is a board which will acquire the next field whether
it is the even or odd field. I understand the early image grabber boards did
work this way. My problem is that I need to acquire an image which is
illuminated by a single srobe flash (on the order of 10 nsec duration). Both
Illumination and acquisition are triggered from an external source which is not
controlled. I've tried synchronizing the strobe to flash twice (once at each
video field), but this results in to much motion artifact. If I grab and
display an entire video frame, I get an excessive amount of flicker. If I grab
an entire frame but display only one field, about half of the time I display
non-illuminated images. If I wait for the next even or odd field, I can have up
to a 30 msec delay, which is unacceptable.

Any information about video boards or other solutions to this problem will be
appreciated. At this point I'm open to almost any suggestion. Thanks.


Dave DeFily defily-at-tamu.edu
Medical Physiology, Texas A&M



-Dave defily-at-tamu.edu





From: CSENCSITS-at-AAEM.AMC.ANL.GOV
Date: Mon, 21 Nov 1994 11:09:02 -0600 (CST)
Subject: Re: Dave Calverts "jokes"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Not appropriate, and not appreciated!
Roseann




From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Mon, 21 Nov 1994 14:27:01 -0500
Subject: "jokes"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There is no room on the microscopy forum for crude jokes of
this type. It is, in my opinion, very unprofessional to send jokes of
this type across a public forum of this calibre. We should try to keep
all information and questions related to microscopy and not to jokes that
will offend a lot of individuals.
Let's try to act professional from now on!
Phil 8-{(




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Mon, 21 Nov 1994 14:44:21 -0600 (CST)
Subject: In apropriateness...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sorry Folks,

I didnot find the joke posted by Calvert either funny or appropriate.
He is now history and has been removed from the listserver. I do not
want to start a moderated system because I don't have the time, however
I will not hesitate to stomp on things like that.

Nestor
Microscopy SysOp




From: Tony D Perdue :      tdperdue-at-email.unc.edu
Date: Mon, 21 Nov 1994 14:53:15 -0500 (EST)
Subject: Re: Dave Calvert's comments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Perhaps Mr. Calvert should have spent more time and energy composing his
"real message".

Perhaps Mr. Calvert simply typed the wrong address for this posting.

Perhaps I have subscribed to the wrong list...

Perhaps this particular brand of "humor" with blatant political/sexist
content seems out of place and is unwelcomed by other subscribers as
well.





From: peling-at-amnh.org (Peling Melville - Interdepartmental Facilities)
Date: Mon, 21 Nov 1994 14:52:34 -0500
Subject: Re: Dave Calverts "jokes"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199411211949.OAA04545-at-science.amnh.org}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I also agree that those jokes ARE NOT appropriate for this server. They
had absolutely nothing to do with microscopy to say the least.

Peling

--------------------------------------------------------------
Peling Melville peling-at-amnh.org
Interdepartmental Laboratories
American Museum of Natural History






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 21 Nov 1994 14:25:11 -0800 (PST)
Subject: Re: LM - Leitz Laborlux info?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Leitz is now a division of Leica. Try (800)716-0800

Regards,

Glen MacDonald
Hearing Development Laboratories RL-30
University of Washington
Seattle, WA 98195
(206)543-8360
glenmac-at-u.washington.edu






From: A. Kent Christensen :      akc-at-umich.edu
Date: Mon, 21 Nov 1994 18:18:52 -0500 (EST)
Subject: Re: LM - Leitz Laborlux info?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Several years ago, Leitz was purchased by Cambridge, which in turn
was purchased by someone else, etc. The company that includes what was
Leitz is now called Leica. An address from 1992 is: Leica Inc., 111 Deer
Lake Road, Deerfield, Illinois 60015, Tel: 708-405-0123, Fax:
708-405-0147. The world headquarters for microscopes is (again, from
some time ago): Leica Mikroskopie und Systeme GmbH, Ernst Leitz-Strasse,
P.O. Box 2040, W 6330 Wetzlar 1, Tel: 64 61-29-0, Fax: 64 41-29-33 99.
You can probably find a more convenient source in the Washington D.C. area.

--------------------------------

On Mon, 21 Nov 1994, Mark O. Walderhaug wrote:

} Gentle List Readers,
} I have recently been given a Leitz Laborlux 11 Microscope.
} It presently only has one objective lens and I'm interested
} populating the turret with more objectives. I'm having a
} hard time finding any information about Leitz. Am I looking
} in the wrong place, or has it been swallowed up by another
} manufacturer?
} Would some kind person point me in the right direction to
} find out about getting objectives for this microscope?
} Thank you for you patience and kindness.
}
} Mark O. Walderhaug voice: 202 205-4682 fax: 202 401-7740
} Microbial Ecology Branch HFS-517 Food and Drug Administration
} 200 C St. S.W., Washington, DC 20204 USA BITNET: MOW-at-BFD
} Internet: mow-at-vm.cfsan.fda.gov or mow-at-bfd.ssw.dhhs.gov
}




From: tayloe-at-rorc.usbm.gov
Date: Mon, 21 Nov 1994 17:55:21 -0600 (CST)
Subject: Re: LM - Leitz Laborlux info?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Mon, 21 Nov 1994, Mark O. Walderhaug wrote:
} Gentle List Readers,
} I have recently been given a Leitz Laborlux 11 Microscope.
} It presently only has one objective lens and I'm interested
} populating the turret with more objectives. I'm having a
} hard time finding any information about Leitz. Am I looking
} in the wrong place, or has it been swallowed up by another
} manufacturer?
} Would some kind person point me in the right direction to
} find out about getting objectives for this microscope?
} Thank you for you patience and kindness.

Gentle List Poster,

Altho' I am not entirely too kind, I will answer your question...
Yes, Leitz has been swallowed up "buy" another. Cambridge Instruments
and Wild Leitz "merged" in April of 1990. Their new name, at least in
the US, is Leica. You may have luck at, according to the letter they
sent me:

Leica Inc.
111 Deer Lake Road
Deerfield, IL 60015 [may have changed to 60090??? sorry, don't know...]
(708) 405-0123
fax: (708) 405-0147 [or: (708) 405-0030???]

The different info within the [..] are from my American Laboratory
(February 1994) Buyers' Guide. If anyone desires to subscribe (free to
qualified individuals) to American Laboratory, they can be reached at:

ISC, Inc
30 Controls Drive
PO Box 870
Shelton, CT 06484-0870
(203) 926-9300
fax: (203) 926-9310

Question: is this Laborlux 11 you received a newer or older model of the
Laborlux S [1989 model]? Just curious...

You are quite welcome, and hope this helps you.
-Rob
,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,,
| Rob Tayloe | MSM Spelunkers Club /\v/\ |
| Metallographic Lab. | Missouri Speleological Survey /\v/\ |
| Rolla Research Center | Bat Conservation International /\v/\ |
| U.S. Bureau of Mines | Missouri Cave & Karst Conservancy |
| tayloe-at-rorc.usbm.gov | National Speleological Society #32993 /\v/\ |
| (314) 364-3169 x247 | American Cave Conservation Association |
''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''





From: marilyn-at-cemmsa.adelaide.edu.au (Marilyn Henderson)
Date: Tue, 22 Nov 1994 10:21:01 +0900
Subject: Reply-In Situ Experiments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199411220048.AAA18539-at-traminer.cemmsa.adelaide.edu.au}
X-Sender: marilyn-at-traminer.cemmsa.adelaide.edu.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Randi,

Two years ago I was fortunate enough to attend the International Conference
on Current Trends in Immunomicroscopy at the George Washington University
Medical Center, where Dr.Gwen Childs gave a great talk on In Situ
Hybridization. At that time her email address was "GCHILDS-at-UTMBEACH". I'm
sure she'd be of help. References at that time were:

Childs et al, Molecular Endocrinology 1:926-932 (1987).
Hoeffler et al, Histochem J 18:597-604 (1986).
Childs et al, Endocrinology 130:335-344 (1992).
" " " " 131: (July,1992)






From: Benyam.Estifanos-at-geol.lu.se
Date: Tue, 22 Nov 1994 09:25:47 +0100
Subject: Re: In Situ Experiments

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Randi!

You might need to read this book:
"Dynamic Experiments In The Electron Microscope",E.P.Butler,K.F.Hale;
Practical Methods in Electron Microscopy Vol 9 (ed. A.M.Glauert);
North-Holland P.C.;1981.
In general, in situ experiments could reveal oriented phase transition
(topotaxy....) and structural disorder (domain, superstructure, modulation..).

Good luck!

Benyam




From: Felix De La IgIesia 7349 :      delaigf-at-aa.wl.com
Date: Tue, 22 Nov 1994 07:15:55 -0400 (EDT)
Subject: UNSUSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta PETVAX.MUAS; Relayed; Tue, 22 Nov 1994 07:15:55 -0400
Mr-Received: by mta PETVAX; Relayed; Tue, 22 Nov 1994 07:15:56 -0400
Mr-Received: by mta SRVR01; Relayed; Tue, 22 Nov 1994 07:16:57 -0400
Disclose-Recipients: prohibited

UNSUSCRIBE MICROSCOPY delaigf-at-mr.research.aa.wl.com






From: sking-at-eleceng.ucl.ac.uk (Simon King)
Date: Tue, 22 Nov 1994 13:12:16 +0100
Subject: Re: Dave Calvert, in-situ Xpts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {24134.9411221312-at-busby}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Perhaps Mr. Calvert should have spent more time and energy composing his
} "real message".
}
} Perhaps Mr. Calvert simply typed the wrong address for this posting.

Another option might be that he left his machine unattended and it wasn't
him at all. Applications such as Eudora do not require a password to _send_
messages (well, it doesn't on our system, anyway) and, theoretically,
_anyone_ could send messages purporting to be me (or anyone else).

Thankfully, we don't seem to have anyone that puerile here :-)

If it was indeed an attempt at humour, then it is to be condemned, but the
whole thing seems so _unlikely_ that I, at least, am not jumping to hasty
conclusions.

Anyway, may as well add another tuppence worth while I'm here...

Concerning in-situ experiments: One suggestion might be to contact Gatan
(say) for a brochure of available TEM stages. This may provide a range of
heating, cooling, straining, whatever stages with suggested and actual
experiments. There are innumerable heating/cooling/straining studies out
there.

Slightly more off the beaten track these days - considerable radiation
damage research has been performed in-situ: One of the most elegant (and
difficult!) being by TJ Black, ML Jenkins, MA Kirk (and possibly IM
Robertson, CA English ?) in the early 1980's in which disordered zones
created by cascade collapse during ion bombardment were imaged in
weak(ish)-beam dark-field using a superlattice reflection of
initially-ordered Cu3Au in-situ in the Argonne Tandem/HVEM. This was done
using a liquid He stage.

Ahh.. found these refs..
TI: COLLAPSE OF DEFECT CASCADES TO DISLOCATION LOOPS.
AU: Kirk_MA Robertson_IM Jenkins_ML English_CA Black_TJ Vetrano_JS
JN: Journal of Nuclear Materials 1987 Vol.149 No.1 pp.21-28

TI: DISPLACEMENT CASCADE COLLAPSE IN Cu//3Au AT LOW TEMPERATURES.
AU: Black_TJ Jenkins_ML Kirk_MA
JN: Institute of Physics Conference Series 1984 No.68 pp.343-346

There will be earlier ones, but BIDS doesn't go back before 1983 and I
don't have the relevent file to hand. See the above for refs. I seem to
remember there was a Proc. Roy. Soc. article on this (?), and it wouldn't
surprise me if there was something in Phil. Mag. A as well...

Hope this helps - you certainly have a large field to cover/select from!



--------------------------------------------------------------------------
Dr. S.L. King
Dept. Electronics and Electrical Engineering,
University College London
Torrington Place,
London WC1E 7JE
England

Tel. : (+44) 171 387 7050 x 3196
Fax. : (+44) 171 387 4350
Email: sking-at-eleceng.ucl.ac.uk






From: gt5307c-at-prism.gatech.edu
Date: Tue, 22 Nov 1994 08:35:43 -0500
Subject: Re: Dave Calvert, in-situ Xpts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: Daniel L. Callahan :      dlc-at-owlnet.rice.edu
Date: Tue, 22 Nov 1994 08:25:50 -0600 (CST)
Subject: Aluminum substrate dissolution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We are attempting to remove the aluminum disk from beneath the magnetic
layers on a hard disk. Given the problems with polishing aluminum
mechanically, the decision was made to try chemical methods first. I
realize that there are plenty of sources on chemical thinning and
electropolishing, but what I'm looking for is good advice on what will
work readily to eliminate 100 microns to 1 mm of alloyed aluminum and
leave the NiP and higher layers alone. We have tried heated gallium with
no great success (perhaps too low of Al solubility for our mass). Any
assistance would be appreciated.


Daniel L. Callahan
Department of Mechanical Engg. and Materials Science
Rice University
dlc-at-owlnet.rice.edu




From: sjames-at-lamar.ColoState.EDU (Sue James)
Date: Tue, 22 Nov 1994 07:52:21 -0700
Subject: Aluminum substrate dissolution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411221450.AA255758-at-lamar.ColoState.EDU}

unsubscribe





From: Tim Foecke :      tfoecke-at-NIST.GOV
Date: Tue, 22 Nov 1994 10:57:16 -0500
Subject: Sony UP-D7000 printer driver for Mac

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Help! We just bought a Sony UPD-7000 dye-sub
printer for our computer facility, and don't
have the printer driver for the Mac. Does
anyone have it and can share, if it's not
illegal, or have any suggestions as to where
to get it? I've tried all the Sony reps I
can find, to no avail. We have the printer, so
I think we would have the rights to the driver
as well. I'm interested in either a Chooser
document or a Photoshop export module. thanks

Tim Foecke
NIST

tfoecke-at-nist.gov





From: Fermin, Cesar :      Fermin.Cesar-at-tmc.tulane.edu
Date: 22 Nov 1994 09:40:01 -0600
Subject: in situ-hybrid. or materials?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



in situ-hybrid. or materials?

Judging from the answers posted, it is not clear what type of in situ
experiments are in question. I personally posted a book title on immuno-in
situ directly to the user asking question. Please be specific when theme is
ambigous. Never assume that only users of your own background are reading the
message! Thanks.

************************************************************
*Cesar D. Fermin, Ph.D \|*|/ Fax (504) 587-7389 *
*Tulane Medical School /|*|\ Answ. Mach.(504) 584-261 *
*Pathology/SL79 \|*|/ Secretary (504) 584-2436 *
*New Orleans, La 70 112 /|*|\ Lab (504) 5841 *
*Fermin-at-TMC.Tulane.edu -} Director of Morphological Services*
************************************************************




From: Lisa Detter-Hoskin :      lisa.detterhoskin-at-gtri.gatech.edu
Date: Tue, 22 Nov 1994 13:23:51 -0600 (CST)
Subject: RE: Edge Wick Experiments of a Paper Laminate Material

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English

Hello Fellow Microscopists,


I am looking for some advice on methodology to use to study the internal
structure of a paperboard laminate which will undergo edge wick, permeation
experiments. In this experiment, four fresh cuts are made on a square
sheet of a paper laminate which is then placed into an aqueous bath.
Theoretically, the fluid should be equally absorbed by the laminate
assuming its fiber matrix is randomly oriented -- this does not seem to be
the case with one of our samples. We must identify why this one sample
performs differntly.

We will use microtomy to prepare cross-sections of this laminate and our
FEG SEM to evaluate its internal structural orientation. From my experience
working with paper samples, I know the visual artifacts may be subtle
without the help of a staining system. Does anyone have any recommendation
of a (heavy metal/or other type-cellulosic) stain which could
be used to track the degree of permeation either optically or with the the
backscattered electron detector on our SEM?

Any other thoughts about evualting the morphology of our laminate would
be greatly appreciated.


Happy Holidays,


Dr. Lisa Detter-Hoskin
Sr. Research Scientist
Materials Analysis Center
Georgia Tech Research Institute
Atlanta, GA 30332-0827

PHONE: (404)894-3460
FAX: (404)894-6199




From: kayton-at-ohsu.edu (Robert Kayton,MAC,CROET)
Date: Tue Nov 22 12:12:01 PST 1994
Subject: Tiny bubbles in the spurrs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0rA1ZK-0007KIC-at-stjohns.ohsu.edu}
Message-Version: 2
} To: Microscopy-at-AAEM.AMC.ANL.GOV

Has anyone else out there had a problem with little bubbles showing up on
their 0.5 to 1 micron sections from Spurrs blocks? These are very small, and
only appear after I coverslip my slide. The most frustrating characteristic
is that the bubbles are not consistantly present. Suggestions are
appreciated.

E-mail address kayton-at-ohsu.edu




From: tivol-at-tethys.ph.albany.edu
Date: Tue, 22 Nov 1994 15:18:24 EST
Subject: Aluminum substrate dissolution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Daniel,
Since aluminum is complexed by OH-, a mild base may do the job. I'm
not sure what is alloyed with the aluminum on a hard disk, but maybe it could
be complexed as well--if it is magnesium, a phosphate will probably do the job.
Good luck.
Yours,
Bill Tivol




From: sjames-at-lamar.ColoState.EDU (Sue James)
Date: Tue, 22 Nov 1994 16:02:21 -0700
Subject: Aluminum substrate dissolution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411222300.AA122491-at-lamar.ColoState.EDU}

unsubscribe sjames-at-lamar.colostate.edu







From: sjames-at-lamar.ColoState.EDU (Sue James)
Date: Tue, 22 Nov 1994 16:02:21 -0700
Subject: Aluminum substrate dissolution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe sjames-at-lamar.colostate.edu








From: david.wharton-at-stonebow.otago.ac.nz (David Wharton)
Date: Wed, 23 Nov 1994 13:10:23 +1100
Subject: bubbles in Spurr resin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We also have been having major problems with holes appearing in our
ultrathin sections of Spurr resin embedded specimens (nematodes). We're not
sure we've solved this yet but the problem may be due to the resin
dissolving material from some plastic containers or if the blocks are
polymersied uncovered some of the components volatalise off, resulting in
the holes. Has anyone any other suggestions?


____________________
David Wharton
Department of Zoology
P.O. Box 56
Dunedin
Tel (064) (03) 479 7963
Fax (064) (03) 479 7584






From: Lisa Detter-Hoskin :      lisa.detterhoskin-at-gtri.gatech.edu
Date: Tue, 22 Nov 1994 13:28:46 -0600 (CST)
Subject: RE: Edge Wick Experiments of a Paper Laminate Material

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English


Hello Fellow Microscopists,


I am looking for some advice on methodology to use to study the internal
structure of a paperboard laminate which will undergo edge wick, permeation
experiments. In this experiment, four fresh cuts are made on a square
sheet of a paper laminate which is then placed into an aqueous bath.
Theoretically, the fluid should be equally absorbed by the laminate
assuming its fiber matrix is randomly oriented -- this does not seem to be
the case with one of our samples. We must identify why this one sample
performs differntly.

We will use microtomy to prepare cross-sections of this laminate and our
FEG SEM to evaluate its internal structural orientation. From my experience
working with paper samples, I know the visual artifacts may be subtle
without the help of a staining system. Does anyone have any recommendation
of a (heavy metal/or other type-cellulosic) stain which could
be used to track the degree of permeation either optically or with the the
backscattered electron detector on our SEM?

Any other thoughts about evaluating the morphology of our laminate would
be greatly appreciated.


Happy Holidays,


Dr. Lisa Detter-Hoskin
Sr. Research Scientist
Materials Analysis Center
Georgia Tech Research Institute
Atlanta, GA 30332-0827

PHONE: (404)894-3460
FAX: (404)894-6199




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Tue, 22 Nov 1994 13:07:59 -0600 (CST)
Subject: An Apology to the ListServer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: pjj-at-utxvms.cc.utexas.edu (Peter Joyce)
Date: Wed, 23 Nov 1994 08:19:20 -0700
Subject: In-situ (be more specific)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Judging from the answers posted, it is not clear what type of in situ
} experiments are in question. I personally posted a book title on immuno-in
} situ directly to the user asking question. Please be specific when theme is
} ambigous. Never assume that only users of your own background are reading the
} message! Thanks.



Agreed, when I first saw the words in-situ I thought of hot stage
experiments or on-line inspection. Experiments typical in the materials
testing arena or even materials processing. Please provide more detail
regarding in-situ experiments, it is a very broad term. . .

P. Joyce

Peter J. Joyce
Graduate Research Assistant - Materials Science & Engineering
University of Texas at Austin
(512) 471-5723

internet: pjj-at-utxvms.cc.utexas.edu






From: randy nessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Wed, 23 Nov 1994 08:10:38 -0600 (CST)
Subject: Re: bubbles in Spurr resin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


From the other post in this thread, I thought the problem was
coverslip media bubbles. I must not have read it carefully enough. My
best luck for infiltrating specimens has been to leave the specimens in a
50%ETOH/50% Spurrs mixture overnite. The next day, I change to 100%
resin, about four times in eight hours. It doesn't seem to matter how
they are polymerized, in Beem capsules that are open or closed.
Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu



On Wed, 23 Nov 1994, David Wharton wrote:

} We also have been having major problems with holes appearing in our
} ultrathin sections of Spurr resin embedded specimens (nematodes). We're not
} sure we've solved this yet but the problem may be due to the resin
} dissolving material from some plastic containers or if the blocks are
} polymersied uncovered some of the components volatalise off, resulting in
} the holes. Has anyone any other suggestions?
}
}
} ____________________
} David Wharton
} Department of Zoology
} P.O. Box 56
} Dunedin
} Tel (064) (03) 479 7963
} Fax (064) (03) 479 7584
}







From: John Mansfield :      John_Mansfield-at-mse.engin.umich.edu
Date: 23 Nov 1994 10:35:08 -0400
Subject: Re: Coating screens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1426535549.65065-at-mse.engin.umich.edu}

Reply... RE} Coating screens
Reply
from: John Mansfield
North Campus Electron Microbeam Analysis Laboratory
University of Michigan
2455 Hayward
Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX: (313)936-3352
jfmjfm-at-umich.edu or John_Mansfield-at-mse.engin.umich.edu
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html

This response is a little late, but we send our screens to the same people
that JEOL do (or at least did). I believe the company is:
Grant Scientific Corporation Electron Microscopy Products Division 1385 Rock
Island Road Gilbert, South Carolina 29054
803-892-2841
Contact: Dunkelberger, Dana
We have had good luck with them and get our 2000, 4000, and EM420 screens
recoated there.
Hope this helps.






From: nee-at-lanl.gov (Norman Elliott)
Date: Wed, 23 Nov 1994 08:11:29 -0700
Subject: Re: Tiny bubbles in the spurrs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} On Tue Nov 22 12:12:01 PST 1994, Robert Kayton,MAC,CROET wrote:
}
} } Has anyone else out there had a problem with little bubbles showing up on
} } their 0.5 to 1 micron sections from Spurrs blocks? These are very small, =
and
} } only appear after I coverslip my slide. The most frustrating characterist=
ic
} } is that the bubbles are not consistantly present. =7FSuggestions are
} } appreciated.
} }


If there are bubbles in the resin from mixing (likely) it is also possible
to remove them by centrifuging. This avoids the problems of placing
uncured resin in a vacuum system (messy and the hardener generally is more
volatile and you may pump sufficient hardener off to prevent full curing).





Norman Elliott | E-mail: nee-at-lanl.gov
Los Alamos National Lab | Fax: 505-665-2104
MST-7 MS E549 | Voice: 505-667-1587
Los Alamos, NM 87545 |






From: Randi Holmestad :      randih-at-imf.unit.no
Date: Wed, 23 Nov 1994 17:12:10 +0100
Subject: in situ experiments..

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello!

-Thank you all for answers conserning in-situ experiments!

I have my background in materials science, and I guess that was
what my committee thought about when giving me the title
'in situ experiments in the transmission electron microscope',
but it has been interesting to see that 'in situ' can mean different
things (hybridisation) as I also saw in my litterature seach..
I need to mention something about this..

It was exactly the point to find out what people thought about
when seeing the title above.. (I am free to say whatever I want
under the given title, and need to pick out something..)

-Thank you again all of you!

Randi.




From: Dwight Beebe :      beebed-at-ERE.UMontreal.CA
Date: Wed, 23 Nov 1994 11:53:17 -0500 (EST)
Subject: TEM: stigmation, neg. staining, parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


**






From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Wed, 23 Nov 1994 13:18:02 -0400 (EDT)
Subject: Vibration isolation platforms for EMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Due to heavy construction work adjacent to our EM facility, that is
expected last for the next three years, I am looking into the possibility of
installing vibration isolation platforms for 2 or 3 of our instruments. I am
aware of TMC (Technical Manufacturing Corporation, Peabody, MA) that
manufactures such platforms. Are there other manufacturers of vibration
isolation platforms for EMs? Also, I would be grateful to hear comments from
persons who have installed the plaforms in their facilities. Thanks!




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Wed, 23 Nov 1994 13:29:25 -0600 (CST)
Subject: Enough discussion on the Apology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I think we've had enough here. If you want to continue
do it off-line to Dave Calvert at his compuserve address..


Nestor




From: OGUOCHA-at-edison.usask.ca
Date: Wed, 23 Nov 1994 12:38:33 -0600 (CST)
Subject: Help with Info on NIHimage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English

Hello netters,

My friend and I are trying to do simple indexing of some TEM electron
diffraction patterns we obtained from an aluminum alloy. On asking
around, we were refered to this group for info on NIHimage software.
We are graduate students of physical metallurgy. So, if any good soul
out here knows how we can get NIHimage or any other user-friendly and
affordable software to help us complete our work, we shall appreciate
the assistance.

Many thanks in anticipation.

Oguocha

(for Oguocha and Ehab)




From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Wed, 23 Nov 1994 11:43:11 -0500
Subject: Re: bubbles in Spurr resin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Wed, 23 Nov 1994, David Wharton wrote:

} We also have been having major problems with holes appearing in our
} ultrathin sections of Spurr resin embedded specimens (nematodes). We're not
} sure we've solved this yet but the problem may be due to the resin
} dissolving material from some plastic containers or if the blocks are
} polymersied uncovered some of the components volatalise off, resulting in
} the holes. Has anyone any other suggestions?
}
}
} ____________________
} David Wharton
} Department of Zoology
} P.O. Box 56
} Dunedin
} Tel (064) (03) 479 7963
} Fax (064) (03) 479 7584
} Dave:
have you tried vacuum infiltration with Spurrs? On difficult to
infiltrate specimens, I leave the specimen in 3:1(Spurrs:Propylene Oxide)
overnight under vacuum then I leave it in pure resin under vacuum until I
get ready to embed it that afternoon. I then cure it in a vacuum oven
for 8 hours(I use receipe A in the Spurrs handout for the resin).
Hope this helps!
Phil

PS: I leave the caps off of my specimen bottles when I leave them in 3:1
overnight.




From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 23 Nov 1994 16:23:46 -0400
Subject: RE-VacBook/Ordering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subject: Time:2:49 PM
OFFICE MEMO RE:VacBook/Ordering Date:11/23/94

I am sorry to hear that some people have had trouble with delays
in filling their orders for my book 'Vacuum Methods in Electron Microscopy'.
I have checked with the publisher, Portland Press,
and have been assured that orders placed with the following
should be handled promptly:
1. In the U.S.A. and Canada:
Portland Press, Ashgate Publishing Co., Old Post Road,
Brookfield, VT 05036-9704
Ph: 802-276-3162 Fx: 802-276-3837
2. In the UK and Europe:
Portland Press, Commerce Way, Colchester, CO2-8HP, UK
Ph: 0206-796351; Fx: 0206-799331
3. In Japan:
OBK Marketing Services, T's Building 3F,
1-38-11 Matsubara, Setagaya-ku, Tokyo 156
Ph: 03-5300-1658 Fx: 03-5300-1615

I will be very much interested in having comments on the book
from any of you that buy it, to see how well you think I have
succeeded in presenting the information electron microscopists
need about their vacuum systems.
Wilbur C. Bigelow






From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 23 Nov 1994 16:53:24 -0400
Subject: RE-InSituExpts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1426512619.42958-at-mse.engin.umich.edu}

Subject: Time:3:27 PM
OFFICE MEMO RE:InSituExpts Date:11/23/94

Back in the 1960s Professor L. O. Brockway and several of his
graduate students carried out a series of rather elegant
experiments in which they recorded the nucleation and growth
of oxide particles on single-crystal thin films of several
metals in a heated gas reaction stage inside a JEOL JEM-6A
TEM using a cine camera. Here are references to some of
their papers:
J. Electrochem. Soc. Vol 121, No. 11, Nov. 1974, p. 1534 (1974);
ibid, vol. 119, p.899 (1972); "Proc. Symp. Funds. of Gas-Surface
Interact", p. 147, Academic Press,(1967); J. Appl. Phys. Vol 37,
p. 2703 (1966); ibid, Vol 34, p.921 (1963); "Single Crystal Films"
M. H. Francombe & H. Sato, Eds., Macmillan (1964). Good luck
with your talk - hope you pass your exam successfully.





From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Wed, 23 Nov 1994 16:13:59 -0500
Subject: retinas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a good procedure for the fixation, processing and
mounting of whole mammalian retinas?
TIA
Phil




From: Marcelle A Gillott :      magem-at-csd.uwm.edu
Date: Wed, 23 Nov 1994 15:56:23 -0600 (CST)
Subject: Re: Tiny bubbles in the spurrs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Regarding your problem of tiny bubbles in Spurrs.....

You dont say if the bubbles are all over, or only in the blank resin or
the tissue....

- the mixing process can introduce bubbles:
After mixing the resin, centrifuge it - try 5-10 min at high speed on a
table top centifuge using 15 or 50 ml conical tubes, depending on the
volume of resin you are preparing

- If the problem is only within the tissue, try infiltrating under vacuum


good luck

marcelle






From: South Bay Technology :      73531.1344-at-compuserve.com
Date: 23 Nov 94 18:50:26 EST
Subject: Vibration isolation platforms for EMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You may want to contact:

Minus K Technology
11775 Gateway Blvd. #6
Los Angeles, CA 90064

TEL: 310-478-6533
FAX: 310-478-4248

Contact: Dr. David L. Platus

Good Luck!

David Henriks
South Bay Technology, Inc.
1120 Via Callejon
San Clemente, CA 92673

TEL: 800-728-2233
FAX: 714-492-1499





From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 23 Nov 1994 17:19:43 -0400
Subject: RE-TEM Contribs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1426511113.34401-at-mse.engin.umich.edu}

Subject: Time:4:01 PM
OFFICE MEMO RE:TEM Contribs Date:11/23/94

Studies by transmission electron microscopy certainly made a
unique, and extremely important, contribution to elucidating the
microstructural changes that occur during the heat treatment of
steels. This was a matter of fundamental importance in the field
of physical metallurgy, because it served as a foundation for understanding
metallurgical heat treating processes in general,
and lead to the improvement of these processes and the production
of improved metal and alloy products in many instances. Much of
this work was carried out by Subcommittee XI, of Committee E4,
of the ASTM in the early 1950s, and is summarized by a report by
Bill Grube which was published in "Revue Universelle des Mines",
Series 9, Vol. XII, No. 10 (1956) and in the "Proc. of the ASTM"
Vol. 56 (1956). I have a copy of the first of these articles that
I can loan you, if you are interested in seeing it.
Wilbur C. Bigelow (bigelow-at-umich.edu)





From: Michael Rock :      merock-at-u.washington.edu
Date: Wed, 23 Nov 1994 14:08:12 -0800 (PST)
Subject: Re: An Apology to the ListServer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to the apology offered to the listserver by Dave. Apology
accepted, at least by me.

Bummer it had to happen to you, but these sort of fopas are a part of the
system. And we all need to get used to the limitations of this new form
of communication. We all need help SPELLING, and we all say things which
may be offensive (at least to some group in our society), and now that
our communication is being recorded and disseminated via the information
superhighway we better brace ourselves for some high speed cyber-collisions.
After all how many of us know enough about this INTERNET system to be
confident that Dave was the person who sent the jokes? there are a lot of
hidden doors in the NET.

Let's all be a little more understanding of our human counterparts,
someday we may ask for the same.

Last night there was a news story on hackers changing a doctors medical
records, several PAP smear tests were reversed via the INTERNET, some
kind of practical joke (not so funny to those patients "misdiagnosed").

something to think about...

-my two cents worth,

Mike Rock





From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Wed, 23 Nov 1994 14:26:10 -0500
Subject: Re: Vibration isolation platforms for EMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Wed, 23 Nov 1994, M.V. Parthasarathy wrote:

} Due to heavy construction work adjacent to our EM facility, that is
} expected last for the next three years, I am looking into the possibility of
} installing vibration isolation platforms for 2 or 3 of our instruments. I am
} aware of TMC (Technical Manufacturing Corporation, Peabody, MA) that
} manufactures such platforms. Are there other manufacturers of vibration
} isolation platforms for EMs? Also, I would be grateful to hear comments from
} persons who have installed the plaforms in their facilities. Thanks!
}
M.V.:
At one time I had 2 AEI 801 E.M.s and the department decided to redo the
department from the ground floor(where I was) up. I put rubber pads with
a hardness of 80 durometers under the scopes and had no problem with
vibration. They had to jackhammer a pit in a room next to the microscope
and I still could use the scope while they worked. Of course, I did this
only if it was a diagnostic path. case. For routine research using the
EM I used the scope whenever they weren't digging. The noise would drive
you batty. So, I found other uses for my time. This may help or you may
have a condition where you have to have these platforms. I try the least
expensive (the pads were 4"x4" and cost me at the time less than $10.00)
way to do things like this first.
Good luck!
Phil 8-{)




From: Dwight Beebe :      beebed-at-ERE.UMontreal.CA
Date: Wed, 23 Nov 1994 12:16:03 -0500 (EST)
Subject: TEM: stigmation, neg. staining, parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9411231831.AA27759-at-riker.ml.wpafb.af.mil}

Hi, sorry about the blank post: fat fingers. Here's the real one.

Greetings!
=09I am teaching a TEM course this Fall and am having considerable=20
difficulty with two areas:1) the students are finding it very hard to=20
stigmate the objective lens properly, and 2) we are having very poor=20
success with a negative staining protocol using bacteria and T-even=20
bacteriophage. I=D5m hoping that those of you who have taught others to=20
properly align a TEM might send me input on their stigmation=20
technique/instruction. We are using a JEOL JEM 100S, stigmating at 50K=20
(the mini-lens assembly is defective, preventing stigmation at higher=20
mags). The resulting negatives clearly show that the astigmatism has not=
=20
been corrected. I have worked individually with the three students and,=20
while I seem to be able to stigmate the lens correctly, they remain very=20
frustrated by their inability to do so. I now wonder if there are other=20
(better) teaching techniques that I might try. In brief, I stigmate the=20
lens using a holey grid with the beam spread to obtain the most coherent=20
illumination, working at the interface between focus and just=20
over-focus. I find the black fringe inside the hole easier to=20
visualize. Once I find the point of focus, I bring the image to just=20
over-focus and determine which of the two (X,Y) stigmators seems to have=20
the greater effect. Then, one at a time, I adjust the stigmators to=20
yield an even fringe within the hole, as I shift back and forth from=20
focus to just over-focus. I encourage the students to be patient and=20
take their time, and to work with the room and panel lights off, so that=20
the image can be viewed distinctly. Any and all comments will be=20
appreciated.
=09I won=D5t go into detail regarding the negative staining protocol,=20
as I=D5m not sure that it=D5s necessary or relevant to many of the readers.=
=20
However, our problem seems to be that we are not obtaining the negative=20
image that one expects (with either 2% PTA, pH 7, or 2% UA in Bacitracin,=
=20
filtered). We have tried freshly prepared solutions, new bacterial=20
cultures, variations in application of the stains (drop vs. flotation)=20
and still seem to have little success. Anyone out there using these or=20
similar techniques on bacteria or virus? Again, I=D5d appreciate comments=
=20
from all who have insight or experience with these techniques. I have=20
read the appropriate texts and articles and can=D5t seem to pin down the=20
likely source of our problems.
=09Finally, I have been given an Ilford Ilfoprint Mark II Super 12=20
wet print processor and now would like to find a few parts for it. =20
Specifically, chemicals (activator and stabilizer) and the two bottles=20
that function as solution reservoirs. Can=D5t afford to buy the dry to dry=
=20
version. Anybody able to point me to a source? Kodak Polycontrast paper=
=20
is used for the majority of the prints from our darkroom.
=09Thanks for your time and assistance!
=09=09Dwight


Dwight Beebe
IRBV, Dept. de sciences biologiques=09=09beebed-at-ere.umontreal.ca
Universite de Montreal=09=09=09=09Voice:514-872-4563
4101, rue Sherbrooke est=09=09=09FAX:514-872-9406
Montreal, PQ H1X 2B2 Canada







From: Keith R. Hallam :      k.r.hallam-at-uk.ac.bristol
Date: Thu, 24 Nov 1994 09:32:51 +0000 (GMT)
Subject: Re: An Apology to the ListServer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} grammatical, and word-omission errors. Not to beat a man while he is
} down, but there is no reason not to take advantage of one of the greatest
} benefits of writing in a text editor: the chance to view one's work,
} read it over, and modify it if necessary for clarity. This is especially
} true in a public forum.

Don't forget that while some people might be writing brilliant prose on
Windows or Mac machines beautifully ethernetted to the World, others (like me)
are struggling with ancient Unix editors (vi, pico) hoping to get a message
finished and sent off before the software decides to insert spurious characters
or the modem throws a wobbly, kicking us off line, or someone else in the
office picks up the 'phone, inserting a series of odd characters which might, or might not, mean something to the computer, including, perhaps, even posting
something. I hope no-one judges the contents of the message purely on its
presentation - very much a symptom of the eighties and nineties (at least in
this country, it seems).

Keith






From: epicier-at-univ-lyon1.fr (Thierry Epicier)
Date: Thu, 24 Nov 1994 14:39:11 +0100
Subject: "SHRLI-SIMPLY" : TEM and HRTEM simulations on PCs...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everybody,

this message is for an annoucement to microscopists in materials
sciences, concerning a set of programs, "SHRLI-SIMPLY" ((c) T. EPICIER and
M.A. O'KEEFE), running on PCs, and which is available on FTP.UNIV-LYON1.FR,
in PUB/DOS/HRTEM (usual login - anonymous - and binary transfer).
As can be expected from its title, this package is running SHRLI ((c) M.A.
O'Keefe, 1980), which allows HRTEM simulations to be conducted (multislice
method, 128x128 FFT). But is allows more than this : building cells
/supercells (crystalline or 'non-crystalline' nature), drawing/modifiying
structures, calculating 'geometrical' reciprocal lattice sections and/or
dynamical diffraction patterns, helping the user to index S.A.D. patterns,
plotting Pendellosung curves, calculating CTF, numerical diffractograms
(including astigmatism defect), comparing calculated and digitized images,..
Although the programs are running under DOS, a graphical user-interface
makes any operation (calculation, screen display, files selections,
print,...) very easy.
This version of S-SIMPLY is freeware, and has limitations ; however,
all options are correctly (hopefully) working.
Further information is available in the README.TXT and INSTALL.BAT files,
accompanying the SIMPLY1.ZIP and SIMPLY2.ZIP files.
Obviously, feel free to contact me for any comments and/or criticisms
regarding these programs....

P.S. recent publications illustrating (in a very limited way) the use of
S-SIMPLY are :
- "Superlubricity of molybdenum disulphide", J.M. MARTIN et al,
Phys. Rev. B, 48, 14, 10 583
- "Quantitative HREM analysis of potassium incorporation into
cordierite", T. EPICIER et al, Proceed. ICEM 13 (Editions de
Physique, Paris), 1994, p. 391
- "Benefits of HREM for the study of metal-ceramic interfaces", T.E.
and C. ESNOUF, J. Phys. III France, 4, (1994), 1811.
______________________________________________

Thierry EPICIER
GEMPPM-502, INSA de Lyon,
69621 VILLEURBANNE, France
tel : (33) 72 43 84 94 (83 85)
FAX : (33) 72 43 85 28
Email : EPICIER-at-CISMSUN.UNIV-LYON1.FR





From: epicier-at-univ-lyon1.fr (Thierry Epicier)
Date: Thu, 24 Nov 1994 15:24:44 +0100
Subject: "SHRLI-SIMPLY" : TEM and HRTEM simulations on PCs...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everybody,

this message is for an annoucement to microscopists in materials
sciences, concerning a set of programs, "SHRLI-SIMPLY" ((c) T. EPICIER and
M.A. O'KEEFE), running on PCs, and which is available on FTP.UNIV-LYON1.FR,
in PUB/DOS/HRTEM (usual login - anonymous - and binary transfer).
As can be expected from its title, this package is running SHRLI ((c) M.A.
O'Keefe, 1980), which allows HRTEM simulations to be conducted (multislice
method, 128x128 FFT). But is allows more than this : building cells
/supercells (crystalline or 'non-crystalline' nature), drawing/modifiying
structures, calculating 'geometrical' reciprocal lattice sections and/or
dynamical diffraction patterns, helping the user to index S.A.D. patterns,
plotting Pendellosung curves, calculating CTF, numerical diffractograms
(including astigmatism defect), comparing calculated and digitized images,..
Although the programs are running under DOS, a graphical user-interface
makes any operation (calculation, screen display, files selections,
print,...) very easy.
This version of S-SIMPLY is freeware, and has limitations ; however,
all options are correctly (hopefully) working.
Further information is available in the README.TXT and INSTALL.BAT files,
accompanying the SIMPLY1.ZIP and SIMPLY2.ZIP files.
Obviously, feel free to contact me for any comments and/or criticisms
regarding these programs....

P.S. recent publications illustrating (in a very limited way) the use of
S-SIMPLY are :
- "Superlubricity of molybdenum disulphide", J.M. MARTIN et al,
Phys. Rev. B, 48, 14, 10 583
- "Quantitative HREM analysis of potassium incorporation into
cordierite", T. EPICIER et al, Proceed. ICEM 13 (Editions de
Physique, Paris), 1994, p. 391
- "Benefits of HREM for the study of metal-ceramic interfaces", T.E.
and C. ESNOUF, J. Phys. III France, 4, (1994), 1811.
______________________________________________

Thierry EPICIER
GEMPPM-502, INSA de Lyon,
69621 VILLEURBANNE, France
tel : (33) 72 43 84 94 (83 85)
FAX : (33) 72 43 85 28
Email : EPICIER-at-CISMSUN.UNIV-LYON1.FR





From: epicier-at-univ-lyon1.fr (Thierry Epicier)
Date: Thu, 24 Nov 1994 16:12:25 +0100
Subject: SHRLI-SIMPLY : TEM / HRTEM simulations on PCs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everybody,

this message is for an annoucement to microscopists in materials
sciences, concerning a set of programs, "SHRLI-SIMPLY" ((c) T. EPICIER and
M.A. O'KEEFE), running on PCs, and which is available on FTP.UNIV-LYON1.FR,
in PUB/DOS/HRTEM (usual login - anonymous - and binary transfer).
As can be expected from its title, this package is running SHRLI ((c) M.A.
O'Keefe, 1980), which allows HRTEM simulations to be conducted (multislice
method, 128x128 FFT). But is allows more than this : building cells
/supercells (crystalline or 'non-crystalline' nature), drawing/modifiying
structures, calculating 'geometrical' reciprocal lattice sections and/or
dynamical diffraction patterns, helping the user to index S.A.D. patterns,
plotting Pendellosung curves, calculating CTF, numerical diffractograms
(including astigmatism defect), comparing calculated and digitized images,..
Although the programs are running under DOS, a graphical user-interface
makes any operation (calculation, screen display, files selections,
print,...) very easy.
This version of S-SIMPLY is freeware, and has limitations ; however,
all options are correctly (hopefully) working.
Further information is available in the README.TXT and INSTALL.BAT files,
accompanying the SIMPLY1.ZIP and SIMPLY2.ZIP files.
Obviously, feel free to contact me for any comments and/or criticisms
regarding these programs....

P.S. recent publications illustrating (in a very limited way) the use of
S-SIMPLY are :
- "Superlubricity of molybdenum disulphide", J.M. MARTIN et al,
Phys. Rev. B, 48, 14, 10 583
- "Quantitative HREM analysis of potassium incorporation into
cordierite", T. EPICIER et al, Proceed. ICEM 13 (Editions de
Physique, Paris), 1994, p. 391
- "Benefits of HREM for the study of metal-ceramic interfaces", T.E.
and C. ESNOUF, J. Phys. III France, 4, (1994), 1811.
______________________________________________

Thierry EPICIER
GEMPPM-502, INSA de Lyon,
69621 VILLEURBANNE, France
tel : (33) 72 43 84 94 (83 85)
FAX : (33) 72 43 85 28
Email : EPICIER-at-CISMSUN.UNIV-LYON1.FR





From: William R. Oliver :      oliver-at-ipas.afip.mil
Date: Thu, 24 Nov 1994 12:56:38 -0500 (EST)
Subject: Re: An Apology to the ListServer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Wed, 23 Nov 1994, Daniel E. Sampson wrote:

} In response to Dave Calvert's apology, I would to add my two cents about a
} common problem in electronic correspondence: lack of editing...


Also, many mail programs have cancellation commands which, if used
soon enough (e.g. before the letter gets out of the spool), will
allow you to kill the thing before it gets out the door. This won't
save you from "next morning regrets," but it can help you out of
that feeling you get when you hear the door slam and you realize
you've just left your keys in the car...

billo





From: ax567-at-freenet.carleton.ca (Greg McMahon)
Date: Thu, 24 Nov 1994 23:06:43 -0500
Subject: Listserver and Literature Review

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear fellow microscopists and spectroscopists,

A comment was recently posted regarding the appropriatness of the
listserver as an avenue for a "fast literature review". I just
wish to put in my two cents worth on the subject. I think first
of all, a few facts about myself are appropriate. I'm 31 years
old, and recently received my Doctoate in Germany (6 months ago,
to be exact). I am also a real fan of doing proper literature
reviews, and when somebody approaches me with a problem, and for
one reason or another I know they haven't done a thorough
literature review, I get quite upset. This said, I see absolutely
nothing wrong with using the listserver as a means of obtaining
information for such an exercise. After all, I think the main point
behind these comprehensive type examinations is to teach the
student how to make best use of his or her time to solve a
specific problem on short notice. Therefore, I think the use of
the listserver is a very clever idea. It also has obviously
brought whole new insights into in-situ experiments. As a
materials scientist, I would never have come to the idea that
such experiments can also play a significant role in biological
research, as has now been pointed out to me. Furthermore, I
believe that most scientists, as a result of the changing times,
are finding themselves involved in many different projects (upon
completion of their university theses). I think very few of
us, if any, have the time to do complete literature reviews on
all of them, even though all of us would dearly love to such.
A number of years ago, this situation might have been different
(I admit, I can't say for certain), when scientists at
laboratories perhaps had only one or two major projects to
work on.

I think it would be very interesting to hear what
other scientists with more experience think about this. I
would also like to kindly ask Nestor to step in if he believes
this is inappropriate for the listserver.

Greg McMahon
Metals Technology Laboratory
CANMET
Ottawa, Ontario
Canada

E-Mail:
ax567-at-freenet.carleton.ca
vchartra-at-emr.ca





From: epicier-at-univ-lyon1.fr (Thierry Epicier)
Date: Fri, 25 Nov 1994 12:09:47 +0100
Subject: SHRLI-SIMPLY : too much messages. !

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm afraid my recent message about "SHRLI-SIMPLY" was sent 3 or 4 times....
My apologize for that, it was not a marketing strategy !!!
______________________________________________

Thierry EPICIER
GEMPPM-502, INSA de Lyon,
69621 VILLEURBANNE, France
tel : (33) 72 43 84 94 (83 85)
FAX : (33) 72 43 85 28
Email : EPICIER-at-CISMSUN.UNIV-LYON1.FR





From: Evelyn Clausnitzer :      eclausnz-at-eye1.eye.ufl.edu
Date: 26 Nov 1994 12:03:46 EST
Subject: retina processing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {MAILQUEUE-101.941126114645.320-at-eye1.eye.ufl.edu}
To: microscopy-at-aaem.amc.anl.gov

Dear Phil:
We process retinas for histology, TEM, and SEM: frog, chicken, mouse, rat,
rabbit, pig, cow and human. We alter our processing techniques slightly for
the species and size, and for what we want to end up with (immunolabelling,
EDS, straight morphology, etc.). You are welcome to contact me at:
eclausnz-at-eye1.eye.ufl.edu
Evelyn Clausnitzer
U of Florida
Ophthalmolgy




From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Wed, 23 Nov 1994 16:13:59 -0500
Subject: retinas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {MAILQUEUE-101.941126115206.352-at-eye1.eye.ufl.edu}
To: microscopy-at-aaem.amc.anl.gov


Does anyone have a good procedure for the fixation, processing and
mounting of whole mammalian retinas?
TIA
Phil

Dear Phil:
We process retinas of several species and for a variety of purposes. Please
contact me at eclausnz-at-eye1.eye.ufl.edu for details.
Evelyn Clausnitzer
U of Florida
Ophthalmolgy




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sun, 27 Nov 1994 14:27:30 -0600 (CST)
Subject: Spare Parts: Philips 501

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Fellow Subscribers...

I'm looking for anyone that has parts for a
Philips 500 series SEM system. I'm in the process of modifying
the secondary electron detector system and
am looking for as much as possible of a second
SEI detector system (grid, scintillator, pmt...)
to save the cost of building a complete new detector.
If you've got any of these items and no longer need
them can you please touch base with me off-line.


Thanks in Advance...

Nestor J. Zaluzec

Email: Zaluzec-at-AAEM.AMC.ANL.GOV
================================




From: m.dickson-at-unsw.edu.au (Melvyn R. Dickson)
Date: Mon, 28 Nov 1994 10:30:25
Subject: Re: TEM: stigmation, neg. staining, parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-aaem.amc.anl.gov


} Subject: TEM: stigmation, neg. staining, parts

} Dwight Beebe
} IRBV, Dept. de sciences biologiques=09=09beebed-at-ere.umontreal.ca
} Universite de Montreal=09=09=09=09Voice:514-872-4563
} 4101, rue Sherbrooke est=09=09=09FAX:514-872-9406
} Montreal, PQ H1X 2B2 Canada

Hi Dwight. We have a lot of experience in teaching students both microscopy
and preparation. First can I suggest you think what is most important for the
class you are taking. Virus morphology, preparation or microscope operation?
For our virology class, we make a very heavy concentration of T4 phage with
a sucrose density gradient centifugation so they deal with a suspension known
to have ample virions. (And we check it before hand). We use 1 p.c PTA with
bacitracin. I use parlodion filmed grids coated with carbon, then exposed to
ion discharge within a few hours of use to guarantee wettability. The
students apply the phage suspension to the filmed grids and allow adsorption
for say 5 minutes. Then they wash away the sucrose by touching the grid on
three drops of distilled water ( 1% sodium acetate if you want to add some
ions). Then they add a drop of PTA. For best negative staining they should
blot off the stain almost at once as with prolonged contact you get some
positive staining as well.

We let each student put in one of their grids and look around. We try to let
them experience some of the excitement of seeing something they don't
understand first, then understand it by say turning up the magnification or
finding a virion. They get prompted with questions like "what do you see?
describe it to us? what do you think it is? etc. They take a picture (or two)
each. I don't bother getting them to do astigmatism as its only a 4 hour
class (all the microscopy you can learn in 4 hours!).

I only teach astigmatism correction when the students are intended to become
independent operators. This occurs in one on one coaching in microscope use.
I TELL students about astigmatism in introductory lectures but I dont expect
them to absorb much except the name and that its an ever attendant problem
while focussing. In fact I introduce the stigmators as being additional focus
controls for two cylindrical lenses to correct defects in the main lens. I
use the model of human astigmatism being corrected with additional lenses.

I gave away using holes for correction about 25 years ago. Even moving from
one hole to the next on a grid can change the environment enough to introduce
a little astigmatism and I assert that changing grids is not acceptable.
Putting in a separate grid was all right when you moved the aperture by
thumping the column with a rubber hammer, but not now.

I teach astigmatism correction by observing the structure of the phase
granularity you can see near focus. For REALLY GROSS astigmatism (most often
introduced by someone giving the knobs a good old wind) I use a bit of the
edge of the grid and correct on the fringes there. A TV system on the column
is great for demonstrating these effects.

For phase granularity correction; wind the objective focus to and fro and
watch the background structure. If it goes from streaks one way to streaks
the other, there is astigmatism. Find the objective focus that gives the
"best" focus. Use ONE astigmatism control like a focus control; wind it to
and fro and find the "best" focus. Repeat with the other astigmatism control.
Repeat the whole cycle, starting with the objective until the background
structure no longer makes streaks, but makes circles that vary in size as the
focus is changed. With experience you can make nearly perfect correction and
even recognise a required amount of defocus. At real focus the granularity
becomes vanishingly small (phase contrast at a minimum) and everything looks a
little fuzzy.

Please send feed back on how useful you find these ideas.

all the best in casting microscopic pearls of wisdom


mel dickson




From: EMLAB-at-opus.mco.edu
Date: Mon, 28 Nov 1994 08:54:21 -0400 (EDT)
Subject: Re: TEM: stigmation, neg. staining, parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dwight,

Practice makes perfect, that is the only recommendation I have for astigmating
the microscope.
In my many years, I have always prefered using a 1% ammonium molybdate
solution for negative staining. It is not a dense as PTA or UA but it seems
to be a better "overall" stain.
What are your samples in? If in a sucrose gradient (for virus) the sucrose
interfers with the negative staining, also are your bacteria in a growth media?
After adhereing the samples to a carbon-coated grid, try a brief rinse in
dH2O, than stain for 10-15 sec.
If lack of sample on grids is the problem, try using a glow discharged grid.
No help for you with the print processor parts.

Good Luck,

Ed Calomeni






From: Lata Prabhu 512-356-7894 :      Lata.Prabhu-at-SEMATECH.Org
Date: Mon, 28 Nov 1994 09:02:00 -0600 (CST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

MR-Received: by mta GATEV3; Relayed; Mon, 28 Nov 1994 09:26:24 -0600 (CST)
Alternate-recipient: prohibited
Disclose-recipients: prohibited


Please unsubscribe lata-prabhu-at-sematech.org
Lata







From: hunt-at-msc.cornell.edu
Date: Mon, 28 Nov 1994 11:40:58 -0500 (EST)
Subject: Re: InGap on GaAs (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Reply-To: hunt-at-msc.cornell.edu

Forwarded message:
} From dast-at-msc.cornell.edu Wed Nov 23 18:56:00 1994
} Reply-To: dast-at-msc.cornell.edu
} From: dast-at-msc.cornell.edu
} Message-Id: {199411232355.AA11766-at-jordi.msc.cornell.edu}
} X-Originated-From: jordi.msc.cornell.edu
} X-Msc-Version: IDA Client - Sparc
} Subject: Re: InGap on GaAs (fwd)
} To: hunt-at-msc.cornell.edu
} Date: Wed, 23 Nov 1994 18:55:49 +0000 (GMT)
} Cc: dieguez-at-iris1.fae.ub.es
} In-Reply-To: {199411231425.AA25253-at-borg.msc.cornell.edu} from "hunt-at-msc.cornell.edu" at Nov 23, 94 09:25:08 am
} X-Mailer: ELM [version 2.4 PL21]
} Mime-Version: 1.0
} Content-Type: text/plain; charset=US-ASCII
} Content-Transfer-Encoding: 7bit
} Content-Length: 2130
}
} Angel,
}
} See comments below - in CAPS
}
}
}
} } Forwarded message:
} } From vaytek-at-INS.INFONET.NET Sat Nov 19 05:53:29 1994
} X-Originated-From: lynx.msc.cornell.edu
} X-Msc-Version: IDA Server
} Date: Fri, 18 Nov 1994 16:47:23 CST
} From: vaytek-at-INS.INFONET.NET
} To: microscopy-at-aaem.amc.anl.gov
} Message-Id: {00987A8C.867A916E.336-at-INS.INFONET.NET}
} Subject: InGap on GaAs
}
} From: MX%"dieguez-at-iris1.fae.ub.es" 16-NOV-1994 18:20:14.00
} To: VAYTEK
} CC:
} Subj:
}
} Return-Path: {dieguez-at-iris1.fae.ub.es}
} Received: from AAEM.AMC.ANL.GOV by INS.INFONET.NET (MX V4.1 AXP) with SMTP;
} Wed, 16 Nov 1994 18:20:11 CST
} Date: Wed, 16 Nov 94 07:48:46 +0100
} From: dieguez-at-iris1.fae.ub.es (Angel Dieguez Barrientos)
} Message-ID: {9411160648.AA09675-at-iris1.fae.ub.es}
} Subsject: CTEM - Need information about lines parallel in a layer
} Apparently-To: microscopy-at-aaem.amc.anl.gov
}
}
} My question is about a group of parallel lines that I have
} observed in a sample of InGaP on GaAs, where the substrate is
} misoriented 10 degrees.
}
} SOUNDS LIKE (001). USUALLY MISORIENTED TOWARDS 110 BUT NEEDS
} TO BE SPECIFIED.
}
} The lines observed appear more clearly
} when I tilt the sample about 30 degrees and I see in the 111
} direction (I am talking about a cross-section). Due that the
} thickness of the layer is 2 microns, I think that the lines are
} not due to moiree fringes provided for the superposition of the
} substrate and the layer. On the other hand, because of the layer
} is clearly observable, I think that the lines not correspond
} to fresnel fringes due to thickness variations.
} If it can help you, this system is ordered by not much (CuPt
} structure). The lines are parallel to the substrate in a 011 cross-
} -section and tilted about 12 degrees in a 0-11 cross-section.
}
}
} COULD IT BE MISFIT DISLOCATIONS ? WHAT IS THE COMPOSITION OF THE
} INGAP ? IS IT A LATTICE MATCHED COMPOSITION ? IF YES, HAS THE
} COMPOSITION BEEN CHECKED WITH SOME OTHER METHOD (RBS, KNOWLEDEGABLE
} X-RAY ? )
}
}
}
}
}
} I need your help!!!!.
} Please help me!!!!!!.
}
} Sinceresly yours A. Dieguez.
}
}
} ieter,
} I thought you might be able to suggest something.
} This is from the microscopy listserver.
} John Hunt
}





From: gkennedy-at-ucsd.edu
Date: Mon, 28 Nov 1994 12:35:05 -0800
Subject: Spurr's bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199411282030.MAA00892-at-ucsd.edu}
X-Sender: gkennedy-at-popmail.ucsd.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
Content-Transfer-Encoding: quoted-printable

I'd be curious to know how many of you that are experiencing problems with=
Spurr's have had these difficulties (or any others) since the additive was=
changed in the NSA of American manufacture. You might remember that I=
requested help with problems using Quetol 651 last year--it appears now=
that the NSA modification is the culprit: I changed to FLUKA NSA and have=
just produced blocks identical to what I had prior to June of 1993. I=
can/will provide anyone interested in details a complete workup and=
explanation as to how I finally solved the mystery. Grace Kennedy UCSD






From: tivol-at-tethys.ph.albany.edu
Date: Mon, 28 Nov 1994 17:42:21 EST
Subject: TEM: stigmation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dwight,
I, too, have trouble with stigmating our HVEM. Some of this comes from
lower contrast at higher voltage, and some is just lack of skill. I have found
it necessary to shoot sets of photos, first for the coarse adjustment (which
usually does not change--unless someone has done something major) and then for
the angle and magnitude of the fine. A significant help has been using our
crude-but-sensitive video system in conjunction with a small condenser aperture
and a large (admits 0.3 nm) objective aperture. Using the small beam gives
high coherence--which you also found useful--but the image is too faint to cor-
rect the astigmatism from the phosphor screen. Using the large obj. aper. im-
proves the contrast without disturbing the fringe (we only get ~0.5 nm resolu-
tion, so a 50 mu aper. is big enough for us; your mileage may vary.). Under
these conditions, I am able to adjust the stigmator *reasonably* well.
Obviously, these procedures might not be appropriate for your instrument and/or
purpose. Good luck; I will be interested in any good ideas which come out of
this discussion.
Yours,
Bill Tivol




From: gkennedy-at-ucsd.edu
Date: Mon, 28 Nov 1994 15:09:09 -0800
Subject: NSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199411282304.PAA08151-at-ucsd.edu}
X-Sender: gkennedy-at-popmail.ucsd.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
Content-Transfer-Encoding: quoted-printable

EMS sent out a letter earlier this year explaining that the formula for the=
additive in NSA which keeps Spurr's blocks light colored has been changed. =
Althought Stacie K. from EMS swears that the change is quite recent, and my=
Quetol problems started in June of 1993, I nevertheless purchased NSA from=
FLUKA--My Quetol problem appears to be completely solved. I just ran 4=
test blocks with two different accelerators with identical results--perfect=
blocks. In the course of the last year and a half, I have tested/run down=
virtually everything you can think of to solve this problem, with no=
progress until now. I was just wondering if the Spurr's bubble problem=
could be related to this change in additive; after all, Spurr's contains a=
component remarkable similar in structure to Quetol. Any ideas???? Grace






From: Richard E. Edelmann :      REDELMAN-at-musom01.mu.wvnet.edu
Date: Tue, 29 Nov 1994 08:32:32 +1100
Subject: SPURR's resin and NSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In addition to Grace Kenndy's comments about NSA, I was wondering
how many people are having trouble with section contamination in the
last year or so. I've personally been having trouble and no of a few
others which have been getting uniform peppering on their sections.
The tourble doesn't seem to be the urnayl acetate, nor lead citrate
(a variety of recepies and batches yield the same results). But
resins using MNA instead of NSA do not seem to have this problem.
Could the new NSA formulation be the source of the problem?


Richard E. Edelmann
Electron Microscopy Facility Supervisor
Marshall University - School of Medicine
Huntington, West Virginia




From: Joyce Craig :      bafpjec-at-uxa.ecn.bgu.edu
Date: Tue, 29 Nov 1994 09:56:12 -0600 (CST)
Subject: Re: TEM: stigmation, neg. staining, parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Teaching them how to stigmate. How ambitious. You seem to be doing the
right thing and have the key words,"time" and "patience". Aren't those
two of the most important concepts in teaching EM, from embedding and
sectioning through photography?

Teaching EM can be frustrating, as it is
sometimes difficult to get students out of the mindset that all they need
do for a course is read, listen, and write. Now they are in a situation
where they have to plan ahead, apply concepts to real situations, and
realize that
the outcome is dependent on hundreds of techniques, chemicals, physical
situations, and equipment conditions, any one of which can be screwed up
with or without the students' knowledge.

It is rewarding when students return (as they occasionally do) and
acknoledge that later success in graduate school, medical school, and
jobs was aided by having taken EM. Some specifically use EM in research
and work, while others no longer use it as a technique, but still have
improved their situations in less tangible ways because of it.




From: DCROMEY-at-CCIT.ARIZONA.EDU
Date: Tue, 29 Nov 1994 10:45:38 -0700 (MST)
Subject: Fixative Quality Control

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In recent weeks I've had occasion to find that two of the labs I work
with have been using formaldehyde stock solutions that were 3-4 years
old. I would like to put together a memo to all the other labs I work
with alerting them to the need to use fixatives of a more "recent"
vintage. I have looked in several histology texts to see if I can find a
guideline for when formaldehyde (either 37% stock, or 10% buffered)
become too old to use, but they haven't been much help. Does someone
have a reference they can refer me to or could you pass on what your
guidelines are?

My experience is primarily in TEM and I know that concentrated
glutaraldehyde tends to polymerize over time and we never let our 3%
glut. exceed 4 months in age. I've learned that formaldehyde gets more
acidic over time, but when is it too far gone?

Thanks for you assistance.


Douglas W. Cromey, M.S.
Cell Biology and Anatomy
Arizona Health Sciences Center
1501 N. Campbell Ave.
Tucson, AZ 85724
(602)626-2824 dcromey-at-ccit.arizona.edu













From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Tue, 29 Nov 1994 13:48:46 -0800 (PST)
Subject: Re: Fixative Quality Control

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer07.u.washington.edu

The same discussion has been going on in our lab with regards to 4%
paraformaldehyde. We have noticed a definite change in as little as one
week after fix preparation, and marked changes after 2 weeks. Our
fixatives are generally kept no longer than 4 weeks. The changes we
notice are that with transcardiac perfusion, the body gets very rigid
within 5 min. when using fix less than 1 week old. It takes longer for
body stifeening after 1 week and athe body remains semi-limp after two
weeks, but the brain histology still looks fine after overnight immersion.
So it might be that there is some diminuition of fixative strength, but
not so much that early postmortem changes are not arrested, and it takes
longer for tho9rough fixation to occur.
Glen MacDonald
Hearing Development Laboratories RL-30
University of Washington
Seattle, WA 98195
(206)543-8360
glenmac-at-u.washington.edu






From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Tue, 29 Nov 1994 19:39:08 -0400 (EDT)
Subject: Formaldehyde/Paraformaldehyde

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Formaldehyde that is prepared from paraformaldehyde powder is normally
made fresh just before use since it very poor shelf life. However, it is a
relatively pure form of formaldehyde and is used usually with glutaraldehyde
as a fixative for EM studies. Commercially available 37% formaldehyde
solution on the other hand has impurities including methanol and therefore
is not good for EM studies. But it can be readily used for studies
at the light microscopy level. It is quite stable compared to the
formaldehyde generated from paraformaldehyde powder. I had the opportunity
a few years ago to compare results obtained from cells fixed with a new and
a 2 year-old stock of the 37% formaldehyde (for the fluorescent-localization
of F-actin with Rhodamine/Phalloidin). I could not detect any difference.




From: kayton-at-ohsu.edu (Robert Kayton,MAC,CROET)
Date: Tue Nov 29 17:12:50 PST 1994
Subject: E.M. Position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0rCdbG-0007KOC-at-stjohns.ohsu.edu}
Message-Version: 2
} To: Microscopy-at-AAEM.AMC.ANL.GOV

RESEARCH ASSISTANT

Faculty Research Assistant Electron Microscopy-1.0 FTE. Requires BS in
biological science (MS preferred),, training in biological lab instrumentation
and protocol. Experience in EM tissue preparation is required. Responsible
for histotechniques. Send Letter, curriculum vitae and names, addresses, and
phone numbers of three references to :

Benita J. Pinz, Executive Assistant
College of Veterinary Medicine
OSU
Corvallis, OR 97331-4801

before January 31, 1995.




From: Greg Erdos ICBR EM Core Lab Univers :      GWERDOS-at-gnv.ifas.ufl.edu
Date: Tue, 29 Nov 1994 21:03:19 -0500 (EST)
Subject: Fixatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Whenever possible we only make enough fixative for that day. That goes for
glut., paraformaldehyde or glut/para mixes. If we need to store it before use
we keep it frozen. We have not done any longevity test but it seems to be good
at least one month. We usually use it up before any longer time. We do the
same with 8% glut vials taht we have opened and not used completely. We seal
the top
with parafilm and then place the vial in a screw top jar with a good seal.

******************************************************
* Greg Erdos ** *
* Director, ICBR EMCL ** Phone 904-392-1295 *
* 218 Carr Hall ** FAX 904-846-0251 *
* University of Florida ** gwerdos-at-gnv.ifas.ufl.edu *
* Gainesville, FL 32611 ** *
******************************************************************




From: norenbur-at-onyx.si.edu (Jon L. Norenburg)
Date: Wed, 30 Nov 1994 09:46:44 -0500
Subject: Re: Fixatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199411300723.CAA05356-at-onyx}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Whenever possible we only make enough fixative for that day. That goes for
} glut., paraformaldehyde or glut/para mixes. If we need to store it before use
} we keep it frozen.
Greg,
You keep the glutaraldehyde frozen? I had understood that that was
inappropriate, but I would be happy to be convinced otherwise. I've not
researched this but I also have never run into a lab where this is done. I
do TEM on a very infrequent basis, but often need to do fixation on very
short notice. So, any improvement in shelf-life of reagents would be a
real boon to me.
Another win for the "net"! :-)
--Jon


Jon L. Norenburg {norenbur-at-onyx.si.edu}
Invertebrate Zoology, National Museum of Natural History
Smithsonian Institution, Washington, DC 20560
Voice 301-238-3508, Fax 301-238-3361






From: Sverker Enestr|m :      sveen-at-pai.liu.se
Date: Wed, 30 Nov 1994 19:23:29 +0100
Subject: Fixative Quality Control

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To members interested in fixatives, formaldehyde solution in particular.

Here are some additional informations about storage of commercial form-
aldehyde.
The principal changes which may take place in formaldehyde on storage are
as follows (listed in their order of importance from a practical standpoint):
(1) Polymerisation and precipitation of polymer.
(2) The Cannizzaro reaction, involving oxidation of one molecule of form-
aldehyde to formic acid and reduction of another to methanol.
(3) Methylal formation.
(4) Oxidation to formic acid.
(5) Condensation to hydroxyaldehydes and sugars.
The changes are detrimental to product quality but may be avoided or kept
at a minimum by maintenance of proper storage conditions. With optimum
conditions of storage, commercial formaldehyde will remain unimpaired for
long periods of time. In general, proper storage involves avoidance of
temperature extremes and the use of storage in glass bottles, inert to
corrosion by the mildly acidic solution. Low temperature favor polymer
precipitation, high temperatures accelerate the reaction leading to
chemical loss of formaldehyde. At improper storage temperatures, a form-
aldehyde solution gradually becomes cloudy and eventually solid hydrated
polymer separates as a precipitate.
Much more could be said about this highly interesting fixative which
possesses many unusual characteristics.
----------------------------------------------
* Sverker Enestr|m *
* Department of Pathology, Link|ping, Sweden *
----------------------------------------------







From: bergrh-at-cc.memphis.edu (R. Howard Berg)
Date: Wed, 30 Nov 1994 14:05:47 +0600
Subject: re: formaldehyde/paraformaldehyde

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since an aqueous solution of formaldehyde (even made fresh from
paraformaldehyde) exists as an equilibrium between methylene glycol and
formaldehyde (skewed heavily towards the m.g.), covalent linkage of
formaldehyde to tissue molecules is very slow , reaching equilibrium in 16
h -at-37 C (eg., reviewed in Fox et al. J Histo Cyto 33:845-853, 1985). This
argues that BRIEF fixation in formaldehyde (as in immuocytochemistry preps)
induces artifacts because of the high concentration of methylene glycol,
and that little actual crosslinking occurs. Why use formaldehyde as a
fixative for EM?

(((((((((((((((((((((((((((((()))))))))))))))))))))))))))))
R. Howard Berg
Biology Department
University of Memphis, Memphis, TN, 38152

phone: 901-678-4449 fax: 901-678-4457
internet: bergrh-at-cc.memphis.edu






From: Paul Webster :      Paul_Webster-at-QuickMail.Yale.edu
Date: 30 Nov 1994 14:00:34 -0400
Subject: Fixatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1425918102.78593-at-QuickMail.Yale.edu}

If you are really interested in reading about fixatives and how they work
there is a chapter devoted to this in the book by G. Griffiths. The ref. is
"Fine Structure Immunocytochemistry" 1993 published by Springer Verlag,
Heidelberg.





From: Alan Pooley :      pooley-at-ahab.rutgers.edu
Date: Wed, 30 Nov 1994 19:53:39 -0500
Subject: Re: Fixative Quality Control

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Let me add a note for those workinf on mineralized tissues after formaldehyde
fixation. Small larval bivalve shells (80 to 250 mis microns) dissiolve
quite ras rapidly in even buffered formaldehyde 10% solution. If fixation
is required in aldehyde, use glutaraldehyde by preference, or use formaldehyde the
minimum time and replace with 70-95% ethanol. or better yet (if only mineralized
material will be studied, use 95% ethanol as sole fixative. Shells last for
years o in ethanol
alan pooley marine sci sem lab rutgers univ