From: STEPHAN HELFER :      S.Helfer-at-rbge.org.uk
Date: Thu, 1 Jun 1995 09:39:03 BST
Subject: Re: SEM: Glutaraldehyde as fixative

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Gail

Having observed powdery mildews in the SEM for considerable time I
have come to three most satisfactory techniques, depending on
whether you would like to observe the conidia or the hyphae on the
leaf surface (in which case the conidia are only in the way and prone
to charging).

1) Obviously the Electroscan SEM would be the ideal instrument for
these observations.

2) Do you have access to CRYO SEM? If not don't dispair; my
favourite technique is to mount fresh specimens (NOT dripping wet!)
onto the stubs and observe them straight in the SEM. You must check
with your technician if this is acceptable ('cause the vacuum pump oil
might get contaminated, however, we found with a turbo pump that
this is not a problem, as long as the procedure isn't done too often
and the backing pump becomes contaminated too quickly).

I normally use a low accelerating voltage (1-5kV) and a short WD with
a reasonably small spot. Using this technique you may observe your
specimens for around 20min before they collaps. The surface
moisture is sufficiently conductive to avert charging and you can go
up to X10,000 without problem.

3) Alternatively, and if the above is unaccepable, I have used EM
grade GA and the standard CPD techniques. You have to be aware,
however, that most of your conidial chains will be disturbed using
chemical fixation (as in fact they will be if you plunge specimens into
liquid nitrogen for CRYO work), and the leaf surface waxes will also
be partially dissolved.

If you are interested I shall email you an image of Rhododendron
powdery mildew generated using option 2) technique above.

Yours sincerely


Dr Stephan Helfer, SSO
Mycologist

Royal Botanic Garden Edinburgh, Inverleith Row, EDINBURGH EH3 5LR,
Scotland UK

email s.helfer-at-rbge.org.uk
phone: +44 (0)131 552 7171 ext 280
or +44 (0)131 459 0446-280 (direct digital VoiceMail line)
fax: +44 (0)131 552 0382




From: Jan Coetzee EM Univ Pretoria :      JANC-at-ccnet.up.ac.za
Date: Thu, 1 Jun 1995 11:23:39 GMT+2
Subject: Re: SEM - powdery mildew

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Regarding the thread on the preparation of delicate fungal structures -

Another possible preparation method that often gives very acceptable
results is Osmium vapour fixation:

Pieces of leaf (or agar) with attached fungal structures are exposed to
the vapours from osmium tetroxide. This is easily done by leaving the
specimens in a closed container (tape sealed petri dish) in the presence of a
few drops of 1 percent aqueous OsO4. Remove the specimens after a day, dry in
a dessiccator over silica gel, mount and sputter coat.
OsO4 vapours are TOXIC, use the fume hood.

Jan Coetzee
Unit for Electron Microscopy Tel:+27-12-420-2075
University of Pretoria Fax:+27-12-342-1738
Pretoria Internet:janc-at-ccnet.up.ac.za
0002 South Africa




From: M. Choudhury :      mc142-at-cus.cam.ac.uk
Date: Thu, 1 Jun 1995 13:39:52 +0100 (BST)
Subject: UNSUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



UNSUBSCRIBE




From: Joe Uknalis :      juknalis-at-arserrc.gov
Date: Thu, 01 Jun 1995 11:09:28 -0400 (EDT)
Subject: MSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

MR-Received: by mta ISHTAR.MUAS; Relayed; Thu, 01 Jun 1995 11:09:28 -0400
MR-Received: by mta ISHTAR; Relayed; Thu, 01 Jun 1995 11:09:29 -0400
Disclose-recipients: prohibited

Sorry folks-
but could some kind soul tell me when/where MSA meeting is this year and
when
abstracts are due.
thanks

Joe

juknalis-at-arserrc.gov






From: sbianchi-at-dbag.unifi.it (Stefano Bianchi)
Date: Wed, 31 May 1995 09:21:44 +0100
Subject: Re: Chain Letter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy {Microscopy-at-aaem.amc.anl.gov}
Message-id: {v01510101abf1d6b0ab34-at-[150.217.45.60]}
MIME-version: 1.0
Content-type: text/plain; charset="us-ascii"
Content-transfer-encoding: 7BIT

At 0:32 31-05-1995, william d.Riley wrote:
} Ladies and or gentleman: PULLeze!! keep such nonsense off this mailing list. I
} believe that most of us don't want to hear it.Nestor? you concur?
?






From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Thu, 1 Jun 1995 16:58:34 +0100 (BST)
Subject: Re: email adres Journal of Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Journal of Microscopy is published in Oxford by Blackwells Science on
behalf of the Royal Microscopical Society. You can iether contact the
Jornal Office by e-mail at :rms-at-vax.ox.ac.uk" or by phone at +44-1865
248768 or by fax +44-1865-791237.
You can contact me General Editor at "pe13-at-cus.cam.ac.uk" or by phone
+44-1223-333946 orfax +-44-1223-333953.
If all this modern and dispassionate forms of communication, it's always
a treat to receive a handwritten letter at Department of Plan Sciences,
University of Cambridge, Downing Street, Cambridge CB2 3EA UK.

Patrick Echlin
June ist 1995
University of Cambridge, Downing Street Cambridge BB2 3EA UK.On Tue, 30 May
1995, Krzysztof Hubner wrote:

}
} Dear Friends
}
} Who know email adres Journal of Microscopy
} (edited in London)
}
} Thanks
} Krzysztof Hubner
}
} Foundry Research Institute
} Krakow, Poland
} {zahubner-at-cyf-kr.edu.pl}
}




From: jeanne_barker-at-merck.com (Jeanne Barker)
Date: 1 Jun 1995 12:33:53 U
Subject: Methyl cellulose solution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506011647.AA08823-at-igw.merck.com}

REGARDING Methyl cellulose solution
Hi,
What is the best way to make a 2% aqueous solution of methyl cellulose?
Should the water be cold, hot or RT? Does it require overnight spinning?
Thanks in advances for your responses.
Jeanne






From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Thu, 1 Jun 1995 12:31:55 -0400 (EDT)
Subject: RE: MSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Thu, 01 Jun 1995 11:09:28 -0400 (EDT),
Joe Uknalis {juknalis-at-arserrc.gov} writes:

} Sorry folks-
} but could some kind soul tell me when/where MSA meeting is this year
} and when
} abstracts are due.
} thanks
}
} Joe
}
} juknalis-at-arserrc.gov
}
---------

Annual meeetings of MSA, MAS and Histochemical Society
will be held jointly at Kansas City Aug 13-17. MSA deadline for papers was
March 15th!

*************************************************************************
M.V. Parthasarathy
Professor of Plant Biology & Director, EM Facility
Section of Plant Biology
228 Plant Science Building
Cornell University, Ithaca, NY 14850
Telephone: (607) 255-1734
Fax: (607) 255-5407
E-mail: mvp2-at-cornell.edu
***********************************************************************




From: Goldmarker-at-aol.com
Date: Thu, 1 Jun 1995 13:43:02 -0400
Subject: Plastoid N..a source!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Would anyone know a source for Plastoid N..an embedding media used for
demineralized bone.

A colleague is attempting to locate a new source for purchase.

Thank you.

Sincerely,

GOLDMARKER-at-AOL.COM




From: Goldmarker-at-aol.com
Date: Thu, 1 Jun 1995 13:43:05 -0400
Subject: Wrinkling acrylic resin..

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A colleague is having difficulty preventing wrinkling of acrylic resin (2-3u
thick) on glass slides. It does not seem to matter whether she uses
pll-coated, gelatin-coated, or uncoated slides.

She was told to add ethylene glycol to her formulation, but I am concerned
about modifying the vendor's formulation.

Does anyone have any suggestions?

Thanks.
Donald P. Cox, Ph.D. [goldmarker-at-aol.com]




From: Elinor Solit :      cambrex-at-world.std.com
Date: Thu, 1 Jun 1995 15:25:51 +0059 (EDT)
Subject: Re: MSA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Joe,

MSA this year is in Kansas City. Abstract due date is a thing of the past.

For full information, suggest you contact Larry Maser or Sharon at the
MSA office, 800-538-3672.

See you in KC.

Ellie Solit
"The Microscope Book"

On Thu, 1 Jun 1995, Joe Uknalis wrote:

} Sorry folks-
} but could some kind soul tell me when/where MSA meeting is this year and
} when
} abstracts are due.
} thanks
}
} Joe
}
} juknalis-at-arserrc.gov
}
}




From: MAILER-DAEMON-at-emout04.mail.aol.com (Mail Delivery Subsystem)
Date: 95-06-01 17:19:07 EDT
Subject: Fwd: #2(2) Returned mail: Host unknown

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



---------------------
Forwarded message:

Mail Split By Gateway

------- cut here --------
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLEASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
------- cut here --------


------- =_aaaaaaaaaa--
------- cut here --------





From: llsutter-at-mtu.edu
Date: 95-05-31 13:48:49 EDT
Subject: Fwd: #10(14) Chain Letter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



---------------------
Forwarded message:

Mail Split By Gateway

------- cut here --------
------- =_aaaaaaaaaa
Content-Type: text/plain; charset="us-ascii"
Content-Description: Session Transcript

550 aaem.amc.gov (tcp)... Host unknown
550 LISTSEVER-at-aaem.amc.gov... Host unknown

------- =_aaaaaaaaaa
Content-Type: message/rfc822
Content-Description: Returned Content

Received: by emout04.mail.aol.com
(1.37.109.11/16.2) id AA250181355; Thu, 1 Jun 1995 17:15:55 -0400

This message came by your list server please send to appropriate party
---------------------
Forwarded message:

Mail Split By Gateway

------- cut here --------
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLEASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLEASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLEASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLEASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE

ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLEASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE ASSHOLE
------- cut here --------





From: ard-at-atom.ansto.gov.au (Arthur Day)
Date: 95-05-31 12:32:36 EDT
Subject: Fwd: Re: Fwd: Chain letter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nice language Art Day
---------------------
Forwarded message:

WELL HERE'S MY 20 MESSAGES, FUCKWIT....




} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 20:16:21 EDT
} From: NJamie
} To: R100186
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 13:17:11 EDT
} From: BZMomOf4
} To: NJamie
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 13:08:57 EDT
} From: BZMomOf4
} To: JessZiG
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 12:38:59 EDT
} From: STEVEJK192
} To: BZMomOf4
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-22 15:57:01 EDT
} From: WHATZUP123
} To: STEVEJK192
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-22 15:23:05 EDT
} From: Lsoccerja
} To: WHATZUP123
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-18 17:43:25 EDT
} From: MPowell840
} To: Lsoccerja
}
}
} ---------------------
} Forwarded message:
} Subj: Chain letter
} Date: 95-05-16 13:55:21 EDT
} From: ALVikings
} To: MPowell840
}
} This message has been sent to you for good luck. The original
} is in New England. It has been sent around the world nine times. The
} luck has now been sent to you. You will receive good luck within four
} days of receiving this message - Provided you, in turn send it on.
} This is no joke. You will receive good luck in your mailbox.
} Send copies to people you think need good luck. Do not keep this
} message.
} This message must leave your hands in 96 hours.
}
} A United States Air Force Officer received 470,000 Dollars.
} Another Man received 40,000 Dollars and lost it because he broke the
} chain.
} Whereas in the Philippines, Gene Welch lost his wife 51 days after
} receiving the message. He failed to circulate the message. However,
} before his death, he received 7,555,000 dollars.
}
} Please send twenty copies and see what happen in four days.
} The chain comes from Venezuela and has written by Saul De Groda,
} A Missionary from South America. Since the copy must tour
} the world, you must make twenty copies and send them to friends and
} associates - After a few days you will get a surprise. This is true,
} even if you are not superstitious.
}
} Do note the following: Constantine Dias received this chain in 1958.
} He asked his secretary to make twenty copies and send them out. A few
} days later he won a lottery of two million dollars. Carlos Daditt, an
} office employee, received the message and forgot that it had to leave
} his
} hands in 96 hours.He lost his job. Later, after finding that message
} again,
} He mailed twenty copies. A few days later he got a better job.
}
} Dalan Fairchild received the message and, not believing - Threw the
} message away. Nine days later he died. In 1987, The message received by
} a young woman in California was very faded and barely readable. She
} promised herself that she would retype the message and send it on, But
} she set it aside to do it later. She was plagued with various problems,
} including expensive car repairs. The letter did not leave her hands
} within
} 96 hours. She finally typed the letter as promised and got a new car.
}
} Good Luck but please remember: 20 copies of this message must leave
} your hands in 96 hours... You must not sign on this message...
}
} --------------
} Forwarded Message:
}
} Date: Wed, May 3, 1995 10:58 PM EDT
} From: LAURA HU
} Subj: Fwd: Very Important
}
} To: ASCOT LOCK, Josh Smith, JAnn Smith, Dev Smith, Josh487, Al Smith,
} Smith Bear, IowaSis, SamRy, SMITH AMY, Mott Smith, HannaSmith, JASSMITH,
} Rudley, Smith WA, Xcgirl, Lis Smith, Cap Scott, Scottkins, ScottT NYC
}
} --------------
} Forwarded Message:
}
} Date: Wed, May 3, 1995 9:16 PM EDT
} From: ByzIIFan
} Subj: Fwd: Very Important
}
} To: Chris
} cc: LiNdA, Larry, Laura, Michael, SHousley, LLCogdell, JLEnterpr, Laura
} KAO, Laura N134, Deb 1063, Laura C F, Laura A H, GMINSD, Laura 512, LAURA
HU,
} Laura Bloo, CHEF LAURA, John, EFX Master, LisaYOG
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 11:22 PM EDT
} From: SEXYLESLEY
} Subj: Fwd: Very Important
}
} To: AmandaC117, RenitaLL, Mpetemoss, Osuwrestl, JERKYBOY27, JAKimble,
} Member7995, Drantie, PPiris, Sarah72981, BECKY121, ByzIIFan, GrandAmlvr,
} Luckynclv, SirAiron, Wiltonier, Serri13, A VANDERPO, BryanAL, Mandie21
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 10:15 PM EDT
} From: Javier237
} Subj: Fwd: Very Important
}
} To: SNOWTIG
} cc: Grrrl99, Peasbi, Kara L Tho, IAmNot4U, ERGARD, Skibuny107, TONI A
} 32, B4Ball, Beth476, Alana618, MEGLN, BARBIE 169, DRD33, Jfezell, Weence,
} SEXYLESLEY, MalCru, Jazmom
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 9:56 PM EDT
} From: Beetle16
} Subj: Fwd: Very Important
}
}
} cc: BCEMT91b, BGBARB, BklynBoy69, Elfcandle, EparagoN, Crmsnprnce, ILE
} MEK, Hot69Rod, Javier237, Jessy ny, LindaB19, MMcke, MRobin7728, PecGuy,
} Piracy77, RSMALL333, Rw44, SMOKEDMAN, TexRob, Urtime
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 2:15 AM EDT
} From: Brooke1209
} Subj: Fwd: Very Important
}
} To: Beetle16
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 30, 1995 1:17 AM EDT
} From: DWatson782
} Subj: Fwd: Very Important
}
} To: BB Dianne, BullBoat, Jenn0024, Beene1, CORIN, Brooke1209,
} JOlson8925, Vmplv, Jene1959, JESSFLA19, Playsoc316, FNicolosi, Axandra,
} Rubie23, Blackdress, Llyr, CATWOMAN17, HAPPY369, HLB810, Pickygrl, MERFIE101
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 29, 1995 2:05 AM EDT
} From: Biehler3
} Subj: Fwd: Very Important
}
} To: DWatson782
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 29, 1995 1:20 AM EDT
} From: Dr Dogg
} Subj: Fwd: Very Important
} To: Eddie B1, Mortalmac2, TheShaka, Araya, BryanB2163, Druboy,
} JYHADMSTR, WILD GAMER, AL8726, OTEROFAM, Cypress69, JerryH7256, Misfit76,
JOE
} GAG2, JoeV455, TRENT69685, Biehler3, JohnnyMK3, CM3001, JodieWeiss
}
} here
}
}
} --------------
} Forwarded Message:
}
} Date: Thu, Apr 27, 1995 9:38 PM EDT
} From: GNelson770
} Subj: Fwd: Very Important
}
} To: Sellback, EGW, Calgal14, Wgaynor, GLO, SGlah, L Lunachic, Gemlover,
} Tink13577, Blueelena, Girly 16, Blond16, Andrea1234, Sage12, BlueEyed G,
} KCousin, DALAN L, Dr Dogg, SuprHST, Skater 1
}
} --------------
} Forwarded Message:
}
} Date: Thu, Apr 27, 1995 8:35 PM EDT
} From: Kevman3379
} Subj: Fwd: Very Important
}
} To: Bob
} cc: RROZZONI, Chrismog, JAC 6688, Dr Dogg, Oreochild, Tim, GNelson770,
} Dr Gamewiz, Lemon, Dekion1, JoelC01, CyanZero, Larson-at-prodigy.com, J
centaur,
} Buuster, FROGLUVER, Bill, Skater 1, Sellback
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 11:48 PM EDT
} From: Jay630
} Subj: Fwd: Very Important
}
} To: SteelDaddy
} cc: Kazin, MogDecker1, Pumpkin288, LemonLemon, MikeEGray, Omega47,
} Sethron, Greensmith, Tre1216, Amon2628, Chocobop, Zed964, Dekion1,
} Kevman3379, Droopy999, AMSLAK, Dazed014, Ladisla, EnsnKim
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 11:42 PM EDT
} From: Mandabarb
} Subj: Fwd: Very Important
}
} To: Jay630
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 6:40 PM EDT
} From: JT Lion
} Subj: Fwd: Very Important
}
} To: Milava
} cc: Mandabarb, MikeRansom, Kestri, Maize69, NuttyPNut, Blonddie, L
} WILKIE, HardKorn, Ketep, DodgersMom, Mysitcal, SWAT MP5, Speclforcs,
} GillianJax, Xaynon, Jasmyne520, RangerBlue, Orc Master, LadyNimue2
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 4:38 PM EDT
} From: Kefka100
} Subj: Fwd: Very Important
}
} To: Ander2099, Sky Ryder3, Delphynus, Miss Imina, Tatiyanna, JT Lion,
} L0rd Knot, Prince Rum, Cyric321, Timon 9, Simba Wi, Bilbo Adin,
VenomChaos,
} Yeyinde, ChewbaccAS, Aquilla, DrgnQueen7, Drythyn, LBite, Tig Irb
}
} --------------
} Forwarded Message:
}
} Date: Tue, Apr 25, 1995 7:35 PM EDT
} From: Jon1Carlo
} Subj: Fwd: Very Important
}
} To: Kefka100
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 24, 1995 11:29 PM EDT
} From: Smutsuqi
} Subj: Fwd: Very Important
}
} To: Librogue, Jon1Carlo, Smutsuqi, MizBird, Pinky1000, NLubin, SonorL,
} idweinberg-at-ucdavis.edu
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 24, 1995 8:06 PM EDT
} From: Pinky1000
} Subj: Fwd: Very Important
}
} To: Jon1Carlo, Smutsuqi, NLubin, DBERLAN, BERLAN, K co Cool, SonorL,
} DanielBarc, Ackbar94, RSQUADRA, KBrewer-at-teetot.acusd.edu,
} EBrewer-at-chaph.usc.edu, VDMF56A-at-prodigy.com, Moten, EL GUIDO, ABJZ,
Happyfoot,
} DrewBald, SNKay, GymShoes
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 22, 1995 6:38 PM EDT
} From: RS Judy XC
} Subj: Fwd: Very Important
}
} To: STUSSY, K co Cool, RSQUADRA, Rebecca674, Tennis121, EBecker789,
} Pinky1000, PUGGY1000, Tails1235, WEETZIE11, LivWorden, FOOTICKLR1, DA
} Tickler, Jymbe, Kimmiebaby, Guide MST, Guide QST, KO Sara, KO Solar
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 22, 1995 6:19 PM EDT
} From: Crickel
} Subj: Fwd: Very Important
}
} To: LilMac6887, Peterfin, B Berreth, RS Judy XC, Ranma11
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 17, 1995 7:43 PM EDT
} From: Eagle516
} Subj: Fwd: Very Important
} To: DAISY11074, Convicts, Sandrasue, Sturf, Crickel, SairaK, ATUZ,
} Alirina, Trackar, Dolby1, Kambot3000, Flyme2nt, SeaDude911, JSCAP, LAMESHA,
} LisaB0069, ECK51, Swing35, Blkhunk, ROBOKELLEN, Eagle516, SMeld26705,
} Jimsatisfy
}
} Send it on
}
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 16, 1995 7:34 PM EDT
} From: SeppL
} Subj: Fwd: Very Important
} To: JGFJr, NEZNARF 15, MacMan0000, Grishka, JFJD, Nirvana340, L Mabius,
} Rangers911, GeoScan, PG Miss, JessyEB, MarninS, LynxV, Groove2, ERIK A9131,
S
} MiLAN, Ardnaxelai, CalNet11, JudithMcK, CARRIE3656, Eagle516, NightieRC,
} ELMER321, MARCUSEE, CORY HUTCH, MaxatCap, JAbrams272, Chido, SimList2, AnjP,
} TQuill2
}
} Here read this
}
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 15, 1995 2:50 PM EDT
} From: DevinP
} Subj: Fwd: Very Important
}
} To: EAEY13A-at-prodigy.com, CARRIE3656, Devster597, Mariah P, DewDewHead,
} NgtvCreep1, NightieRC, ZRT900, SeppL, RAJP-at-vm.temple.edu, TQuill2, S MiLAN,
} RuffAndy, GoldigR, Spfrantz, Hackmast, ELMER321, JHK RLK, Annice, Strs,
} Katers11, Missa100, Randilyn27, Pavlinka19, DanialH, StreetFunk, MGrivet,
} Spiderella, Wolfinator, Mav888, DocMP, TonyP1414, ERBrotman, IS 227,
} Goewyn1018, KAPP54, RENNEA, SkoolKat, Hotvettek
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 12, 1995 10:20 PM EDT
} From: Alyndag
} Subj: Fwd: Very Important
}
} To: DevinP
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 10, 1995 3:40 PM EDT
} From: KRowan1048
} Subj: Fwd: Very Important
}
} To: Alyndag
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 9, 1995 5:11 AM EDT
} From: PACK 15
} Subj: Fwd: Very Important
}
} To: KRowan1048
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 9, 1995 3:42 AM EDT
} From: GROOVY DOG
} Subj: Fwd: Very Important
}
} To: PACK 15
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 8, 1995 3:28 AM EDT
} From: Srotag22
} Subj: Fwd: Very Important
}
} To: JuanjA5463, PBGemini, Crystal914, Thrive95, RGSKMA, GROOVY DOG,
} TLester726, V8thunder, ABOMB23, DncngBare, PJR29, EVEBIME, PsyburMan,
} AdamB51867, JWGASKINS, KitKelly, FRT, CyberLot, BryanOL, Ricky68
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 5, 1995 9:49 PM EDT
} From: Yamahain
} Subj: Fwd: Very Important
}
} To: Srotag22
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 5, 1995 8:15 PM EDT
} From: Punk core
} Subj: Fwd: Very Important
} To: Yamahain, Punker Joe, CleInd, Boyjer
}
} .
}
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 5, 1995 3:15 PM EDT
} From: WarGymnast
} Subj: Fwd: Very Important
}
} cc: Phalanx123, Punk core, Lisashot
}
} here
}
}
} --------------
} Forwarded Message:
}
} Date: Tue, Apr 4, 1995 2:45 PM EDT
} From: Kitt4
} Subj: Fwd: Very Important
}
} To: WarGymnast
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 3, 1995 6:33 PM EDT
} From: ATFRRT
} Subj: Fwd: Very Important
}
} To: Kitt4
}
} --------------
} Forwarded Message:
}
} Date: Tue, Mar 28, 1995 11:10 PM EDT
} From: Payton2
} Subj: Very Important
} To: ATFRRT
} cc: Someone
}
} This message has been sent to you for good luck. The original
} is in New England. It has been sent around the world nine times. The
} luck has now been sent to you. You will receive good luck within four
} days of receiving this message - Provided you, in turn send it on.
} This is no joke. You will receive good luck in your mailbox.
} Send copies to people you think need good luck. Do not keep this
} message.
} This message must leave your hands in 96 hours.
}
} A United States Air Force Officer received 470,000 Dollars.
} Another Man received 40,000 Dollars and lost it because he broke the
} chain.
} Whereas in the Philippines, Gene Welch lost his wife 51 days after
} receiving the message. He failed to circulate the message. However,
} before his death, he received 7,555,000 dollars.
}
} Please send twenty copies and see what happen in four days.
} The chain comes from Venezuela and has written by Saul De Groda,
} A Missionary from South America. Since the copy must tour
} the world, you must make twenty copies and send them to friends and
} associates - After a few days you will get a surprise. This is true,
} even if you are not superstitious.
}
} Do note the following: Constantine Dias received this chain in 1958.
} He asked his secretary to make twenty copies and send them out. A few
} days later he won a lottery of two million dollars. Carlos Daditt, an
} office employee, received the message and forgot that it had to leave
} his
} hands in 96 hours.He lost his job. Later, after finding that message
} again,
} He mailed twenty copies. A few days later he got a better job.
}
} Dalan Fairchild received the message and, not believing - Threw the
} message away. Nine days later he died. In 1987, The message received by
} a young woman in California was very faded and barely readable. She
} promised herself that she would retype the message and send it on, But
} she set it aside to do it later. She was plagued with various problems,
} including expensive car repairs. The letter did not leave her hands
} within
} 96 hours. She finally typed the letter as promised and got a new car.
}
} Good Luck but please remember: 20 copies of this message







From: hukee.margaret-at-mayo.EDU (Marge Hukee)
Date: Thu, 1 Jun 1995 16:30:18 +0200
Subject: Wrinkling acrylic resin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506012128.AA27893-at-fermat.Mayo.EDU}
X-Sender: hukem-at-fermat.mayo.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dr. Cox

We float sections on a water bubble and expose the sections to vapor from
trichloroethylene for a few minutes prior to drying at 55 degrees C. This
usually will flatten sections if the resin has been cured sufficiently. If
you have any questions on this technique, feel free to e-mail.

Marge

Marge Hukee
EM Core Facility hukee.margaret-at-mayo.edu
Mayo Foundation (507) 284-3148
----------------------------------------------------------------------------
--






From: R100186-at-aol.com
Date: Thu, 1 Jun 1995 17:37:00 -0400
Subject: apology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To every one who has not rec a personal apology here it is .I"AM SORRY it was
not intentional I have a new auto address program and (I'am not quite sure
how) it picked up your address and sent that letter. It will not happen
again.





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Thu, 1 Jun 1995 17:22:15 -0500 (CDT)
Subject: Trash Mail from R100186@aol.com

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Subscribers....

Yes I have seen the 2nd trash message. I have contacted
AOL (a second time) to complain about this user. If you wish to add
your complaint. You should address your message to:

Postmaster-at-AOL.COM

let them know you are complaining about the user

R100186-at-aol.com

If AOL sees enough complaints they should take some
action. Because this is an unmoderated list I donot see
the messages until after they have hit the network.
I apologize for this appearing on your screens, but
at the moment there is little I can do.

The best we can do is let AOL know that they have
an individual on their system that is acting in a manner
which is not compatible with a public forum (to say the least).

Nestor





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Thu, 1 Jun 1995 17:30:11 -0500 (CDT)
Subject: Trash Mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Subscribers....


I have send a second notice to AOL about their
user who is dumping junk mail onto the microscopy
listserver.

I notice that he has sent an apology, however, if
it happens again I will let you all know how to
contact AOL to complain. If they get several
thousand complaints then they may actually do
something.

Nestor....

If you do want to contact AOL the generic
address you should use is:

Postmaster-at-AOL.COM

Someone will see it....





From: m.dickson-at-unsw.edu.au (Melvyn R. Dickson)
Date: Fri, 2 Jun 1995 10:26:34
Subject: Micromaze foreline traps

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-aaem.amc.anl.gov

I have been persuaded that micromaze traps are ideal for reducing
backstreaming into microscopes. BUT there are conflicting versions as to how
to run them. The manufacturer says to use them with a valve between the trap
and the (e.g.) microscope and bake the trap out with the valve closed once a
week (or so). I hear of others who just run the trap hot all the time and use
no valve.

Can we have some feedback on the best way to use Micromazes?
Valve?
No valve?
How does it trap if its hot all the time?
Is a simmerstat needed to control operating temperature?

Thanks in advance.

Mel dickson.




From: anncali-at-andromeda.rutgers.edu (Ann Cali)
Date: Thu, 01 Jun 1995 20:44:55 -0400
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506020059.UAA18492-at-andromeda.rutgers.edu}
X-Sender: anncali-at-andromeda.rutgers.edu
X-Mailer: Windows Eudora Version 2.0.3
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

please unsubscribe

Thank you
Ann Cali, Professor (anncali-at- andromeda.rutgers.edu)
Dept of Biological Sciences, Smith Hall
Rutgers University, Newark, NJ 07102
FAX=201-648-1007





From: anncali-at-andromeda.rutgers.edu (Ann Cali)
Date: Thu, 01 Jun 1995 21:04:49 -0400
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506020119.VAA18809-at-andromeda.rutgers.edu}
X-Sender: anncali-at-andromeda.rutgers.edu
X-Mailer: Windows Eudora Version 2.0.3
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


Ann Cali, Professor (anncali-at- andromeda.rutgers.edu)
Dept of Biological Sciences, Smith Hall
Rutgers University, Newark, NJ 07102
FAX=201-648-1007





From: ard-at-atom.ansto.gov.au (Arthur Day)
Date: 95-05-31 03:27:24 EDT
Subject: Fwd: Re: Fwd: Chain letter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {25060120323134-at-vms2.macc.wisc.edu}

Perhaps we are connected to a street gang.
---------------------
Forwarded message:

WELL HERE'S MY 20 MESSAGES, FUCKWIT....




} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 20:16:21 EDT
} From: NJamie
} To: R100186
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 13:17:11 EDT
} From: BZMomOf4
} To: NJamie
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 13:08:57 EDT
} From: BZMomOf4
} To: JessZiG
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-28 12:38:59 EDT
} From: STEVEJK192
} To: BZMomOf4
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-22 15:57:01 EDT
} From: WHATZUP123
} To: STEVEJK192
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-22 15:23:05 EDT
} From: Lsoccerja
} To: WHATZUP123
}
}
} ---------------------
} Forwarded message:
} Subj: Fwd: Chain letter
} Date: 95-05-18 17:43:25 EDT
} From: MPowell840
} To: Lsoccerja
}
}
} ---------------------
} Forwarded message:
} Subj: Chain letter
} Date: 95-05-16 13:55:21 EDT
} From: ALVikings
} To: MPowell840
}
} This message has been sent to you for good luck. The original
} is in New England. It has been sent around the world nine times. The
} luck has now been sent to you. You will receive good luck within four
} days of receiving this message - Provided you, in turn send it on.
} This is no joke. You will receive good luck in your mailbox.
} Send copies to people you think need good luck. Do not keep this
} message.
} This message must leave your hands in 96 hours.
}
} A United States Air Force Officer received 470,000 Dollars.
} Another Man received 40,000 Dollars and lost it because he broke the
} chain.
} Whereas in the Philippines, Gene Welch lost his wife 51 days after
} receiving the message. He failed to circulate the message. However,
} before his death, he received 7,555,000 dollars.
}
} Please send twenty copies and see what happen in four days.
} The chain comes from Venezuela and has written by Saul De Groda,
} A Missionary from South America. Since the copy must tour
} the world, you must make twenty copies and send them to friends and
} associates - After a few days you will get a surprise. This is true,
} even if you are not superstitious.
}
} Do note the following: Constantine Dias received this chain in 1958.
} He asked his secretary to make twenty copies and send them out. A few
} days later he won a lottery of two million dollars. Carlos Daditt, an
} office employee, received the message and forgot that it had to leave
} his
} hands in 96 hours.He lost his job. Later, after finding that message
} again,
} He mailed twenty copies. A few days later he got a better job.
}
} Dalan Fairchild received the message and, not believing - Threw the
} message away. Nine days later he died. In 1987, The message received by
} a young woman in California was very faded and barely readable. She
} promised herself that she would retype the message and send it on, But
} she set it aside to do it later. She was plagued with various problems,
} including expensive car repairs. The letter did not leave her hands
} within
} 96 hours. She finally typed the letter as promised and got a new car.
}
} Good Luck but please remember: 20 copies of this message must leave
} your hands in 96 hours... You must not sign on this message...
}
} --------------
} Forwarded Message:
}
} Date: Wed, May 3, 1995 10:58 PM EDT
} From: LAURA HU
} Subj: Fwd: Very Important
}
} To: ASCOT LOCK, Josh Smith, JAnn Smith, Dev Smith, Josh487, Al Smith,
} Smith Bear, IowaSis, SamRy, SMITH AMY, Mott Smith, HannaSmith, JASSMITH,
} Rudley, Smith WA, Xcgirl, Lis Smith, Cap Scott, Scottkins, ScottT NYC
}
} --------------
} Forwarded Message:
}
} Date: Wed, May 3, 1995 9:16 PM EDT
} From: ByzIIFan
} Subj: Fwd: Very Important
}
} To: Chris
} cc: LiNdA, Larry, Laura, Michael, SHousley, LLCogdell, JLEnterpr, Laura
} KAO, Laura N134, Deb 1063, Laura C F, Laura A H, GMINSD, Laura 512, LAURA
HU,
} Laura Bloo, CHEF LAURA, John, EFX Master, LisaYOG
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 11:22 PM EDT
} From: SEXYLESLEY
} Subj: Fwd: Very Important
}
} To: AmandaC117, RenitaLL, Mpetemoss, Osuwrestl, JERKYBOY27, JAKimble,
} Member7995, Drantie, PPiris, Sarah72981, BECKY121, ByzIIFan, GrandAmlvr,
} Luckynclv, SirAiron, Wiltonier, Serri13, A VANDERPO, BryanAL, Mandie21
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 10:15 PM EDT
} From: Javier237
} Subj: Fwd: Very Important
}
} To: SNOWTIG
} cc: Grrrl99, Peasbi, Kara L Tho, IAmNot4U, ERGARD, Skibuny107, TONI A
} 32, B4Ball, Beth476, Alana618, MEGLN, BARBIE 169, DRD33, Jfezell, Weence,
} SEXYLESLEY, MalCru, Jazmom
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 9:56 PM EDT
} From: Beetle16
} Subj: Fwd: Very Important
}
}
} cc: BCEMT91b, BGBARB, BklynBoy69, Elfcandle, EparagoN, Crmsnprnce, ILE
} MEK, Hot69Rod, Javier237, Jessy ny, LindaB19, MMcke, MRobin7728, PecGuy,
} Piracy77, RSMALL333, Rw44, SMOKEDMAN, TexRob, Urtime
}
} --------------
} Forwarded Message:
}
} Date: Tue, May 2, 1995 2:15 AM EDT
} From: Brooke1209
} Subj: Fwd: Very Important
}
} To: Beetle16
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 30, 1995 1:17 AM EDT
} From: DWatson782
} Subj: Fwd: Very Important
}
} To: BB Dianne, BullBoat, Jenn0024, Beene1, CORIN, Brooke1209,
} JOlson8925, Vmplv, Jene1959, JESSFLA19, Playsoc316, FNicolosi, Axandra,
} Rubie23, Blackdress, Llyr, CATWOMAN17, HAPPY369, HLB810, Pickygrl, MERFIE101
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 29, 1995 2:05 AM EDT
} From: Biehler3
} Subj: Fwd: Very Important
}
} To: DWatson782
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 29, 1995 1:20 AM EDT
} From: Dr Dogg
} Subj: Fwd: Very Important
} To: Eddie B1, Mortalmac2, TheShaka, Araya, BryanB2163, Druboy,
} JYHADMSTR, WILD GAMER, AL8726, OTEROFAM, Cypress69, JerryH7256, Misfit76,
JOE
} GAG2, JoeV455, TRENT69685, Biehler3, JohnnyMK3, CM3001, JodieWeiss
}
} here
}
}
} --------------
} Forwarded Message:
}
} Date: Thu, Apr 27, 1995 9:38 PM EDT
} From: GNelson770
} Subj: Fwd: Very Important
}
} To: Sellback, EGW, Calgal14, Wgaynor, GLO, SGlah, L Lunachic, Gemlover,
} Tink13577, Blueelena, Girly 16, Blond16, Andrea1234, Sage12, BlueEyed G,
} KCousin, DALAN L, Dr Dogg, SuprHST, Skater 1
}
} --------------
} Forwarded Message:
}
} Date: Thu, Apr 27, 1995 8:35 PM EDT
} From: Kevman3379
} Subj: Fwd: Very Important
}
} To: Bob
} cc: RROZZONI, Chrismog, JAC 6688, Dr Dogg, Oreochild, Tim, GNelson770,
} Dr Gamewiz, Lemon, Dekion1, JoelC01, CyanZero, Larson-at-prodigy.com, J
centaur,
} Buuster, FROGLUVER, Bill, Skater 1, Sellback
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 11:48 PM EDT
} From: Jay630
} Subj: Fwd: Very Important
}
} To: SteelDaddy
} cc: Kazin, MogDecker1, Pumpkin288, LemonLemon, MikeEGray, Omega47,
} Sethron, Greensmith, Tre1216, Amon2628, Chocobop, Zed964, Dekion1,
} Kevman3379, Droopy999, AMSLAK, Dazed014, Ladisla, EnsnKim
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 11:42 PM EDT
} From: Mandabarb
} Subj: Fwd: Very Important
}
} To: Jay630
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 6:40 PM EDT
} From: JT Lion
} Subj: Fwd: Very Important
}
} To: Milava
} cc: Mandabarb, MikeRansom, Kestri, Maize69, NuttyPNut, Blonddie, L
} WILKIE, HardKorn, Ketep, DodgersMom, Mysitcal, SWAT MP5, Speclforcs,
} GillianJax, Xaynon, Jasmyne520, RangerBlue, Orc Master, LadyNimue2
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 26, 1995 4:38 PM EDT
} From: Kefka100
} Subj: Fwd: Very Important
}
} To: Ander2099, Sky Ryder3, Delphynus, Miss Imina, Tatiyanna, JT Lion,
} L0rd Knot, Prince Rum, Cyric321, Timon 9, Simba Wi, Bilbo Adin,
VenomChaos,
} Yeyinde, ChewbaccAS, Aquilla, DrgnQueen7, Drythyn, LBite, Tig Irb
}
} --------------
} Forwarded Message:
}
} Date: Tue, Apr 25, 1995 7:35 PM EDT
} From: Jon1Carlo
} Subj: Fwd: Very Important
}
} To: Kefka100
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 24, 1995 11:29 PM EDT
} From: Smutsuqi
} Subj: Fwd: Very Important
}
} To: Librogue, Jon1Carlo, Smutsuqi, MizBird, Pinky1000, NLubin, SonorL,
} idweinberg-at-ucdavis.edu
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 24, 1995 8:06 PM EDT
} From: Pinky1000
} Subj: Fwd: Very Important
}
} To: Jon1Carlo, Smutsuqi, NLubin, DBERLAN, BERLAN, K co Cool, SonorL,
} DanielBarc, Ackbar94, RSQUADRA, KBrewer-at-teetot.acusd.edu,
} EBrewer-at-chaph.usc.edu, VDMF56A-at-prodigy.com, Moten, EL GUIDO, ABJZ,
Happyfoot,
} DrewBald, SNKay, GymShoes
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 22, 1995 6:38 PM EDT
} From: RS Judy XC
} Subj: Fwd: Very Important
}
} To: STUSSY, K co Cool, RSQUADRA, Rebecca674, Tennis121, EBecker789,
} Pinky1000, PUGGY1000, Tails1235, WEETZIE11, LivWorden, FOOTICKLR1, DA
} Tickler, Jymbe, Kimmiebaby, Guide MST, Guide QST, KO Sara, KO Solar
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 22, 1995 6:19 PM EDT
} From: Crickel
} Subj: Fwd: Very Important
}
} To: LilMac6887, Peterfin, B Berreth, RS Judy XC, Ranma11
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 17, 1995 7:43 PM EDT
} From: Eagle516
} Subj: Fwd: Very Important
} To: DAISY11074, Convicts, Sandrasue, Sturf, Crickel, SairaK, ATUZ,
} Alirina, Trackar, Dolby1, Kambot3000, Flyme2nt, SeaDude911, JSCAP, LAMESHA,
} LisaB0069, ECK51, Swing35, Blkhunk, ROBOKELLEN, Eagle516, SMeld26705,
} Jimsatisfy
}
} Send it on
}
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 16, 1995 7:34 PM EDT
} From: SeppL
} Subj: Fwd: Very Important
} To: JGFJr, NEZNARF 15, MacMan0000, Grishka, JFJD, Nirvana340, L Mabius,
} Rangers911, GeoScan, PG Miss, JessyEB, MarninS, LynxV, Groove2, ERIK A9131,
S
} MiLAN, Ardnaxelai, CalNet11, JudithMcK, CARRIE3656, Eagle516, NightieRC,
} ELMER321, MARCUSEE, CORY HUTCH, MaxatCap, JAbrams272, Chido, SimList2, AnjP,
} TQuill2
}
} Here read this
}
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 15, 1995 2:50 PM EDT
} From: DevinP
} Subj: Fwd: Very Important
}
} To: EAEY13A-at-prodigy.com, CARRIE3656, Devster597, Mariah P, DewDewHead,
} NgtvCreep1, NightieRC, ZRT900, SeppL, RAJP-at-vm.temple.edu, TQuill2, S MiLAN,
} RuffAndy, GoldigR, Spfrantz, Hackmast, ELMER321, JHK RLK, Annice, Strs,
} Katers11, Missa100, Randilyn27, Pavlinka19, DanialH, StreetFunk, MGrivet,
} Spiderella, Wolfinator, Mav888, DocMP, TonyP1414, ERBrotman, IS 227,
} Goewyn1018, KAPP54, RENNEA, SkoolKat, Hotvettek
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 12, 1995 10:20 PM EDT
} From: Alyndag
} Subj: Fwd: Very Important
}
} To: DevinP
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 10, 1995 3:40 PM EDT
} From: KRowan1048
} Subj: Fwd: Very Important
}
} To: Alyndag
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 9, 1995 5:11 AM EDT
} From: PACK 15
} Subj: Fwd: Very Important
}
} To: KRowan1048
}
} --------------
} Forwarded Message:
}
} Date: Sun, Apr 9, 1995 3:42 AM EDT
} From: GROOVY DOG
} Subj: Fwd: Very Important
}
} To: PACK 15
}
} --------------
} Forwarded Message:
}
} Date: Sat, Apr 8, 1995 3:28 AM EDT
} From: Srotag22
} Subj: Fwd: Very Important
}
} To: JuanjA5463, PBGemini, Crystal914, Thrive95, RGSKMA, GROOVY DOG,
} TLester726, V8thunder, ABOMB23, DncngBare, PJR29, EVEBIME, PsyburMan,
} AdamB51867, JWGASKINS, KitKelly, FRT, CyberLot, BryanOL, Ricky68
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 5, 1995 9:49 PM EDT
} From: Yamahain
} Subj: Fwd: Very Important
}
} To: Srotag22
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 5, 1995 8:15 PM EDT
} From: Punk core
} Subj: Fwd: Very Important
} To: Yamahain, Punker Joe, CleInd, Boyjer
}
} .
}
}
} --------------
} Forwarded Message:
}
} Date: Wed, Apr 5, 1995 3:15 PM EDT
} From: WarGymnast
} Subj: Fwd: Very Important
}
} cc: Phalanx123, Punk core, Lisashot
}
} here
}
}
} --------------
} Forwarded Message:
}
} Date: Tue, Apr 4, 1995 2:45 PM EDT
} From: Kitt4
} Subj: Fwd: Very Important
}
} To: WarGymnast
}
} --------------
} Forwarded Message:
}
} Date: Mon, Apr 3, 1995 6:33 PM EDT
} From: ATFRRT
} Subj: Fwd: Very Important
}
} To: Kitt4
}
} --------------
} Forwarded Message:
}
} Date: Tue, Mar 28, 1995 11:10 PM EDT
} From: Payton2
} Subj: Very Important
} To: ATFRRT
} cc: Someone
}
} This message has been sent to you for good luck. The original
} is in New England. It has been sent around the world nine times. The
} luck has now been sent to you. You will receive good luck within four
} days of receiving this message - Provided you, in turn send it on.
} This is no joke. You will receive good luck in your mailbox.
} Send copies to people you think need good luck. Do not keep this
} message.
} This message must leave your hands in 96 hours.
}
} A United States Air Force Officer received 470,000 Dollars.
} Another Man received 40,000 Dollars and lost it because he broke the
} chain.
} Whereas in the Philippines, Gene Welch lost his wife 51 days after
} receiving the message. He failed to circulate the message. However,
} before his death, he received 7,555,000 dollars.
}
} Please send twenty copies and see what happen in four days.
} The chain comes from Venezuela and has written by Saul De Groda,
} A Missionary from South America. Since the copy must tour
} the world, you must make twenty copies and send them to friends and
} associates - After a few days you will get a surprise. This is true,
} even if you are not superstitious.
}
} Do note the following: Constantine Dias received this chain in 1958.
} He asked his secretary to make twenty copies and send them out. A few
} days later he won a lottery of two million dollars. Carlos Daditt, an
} office employee, received the message and forgot that it had to leave
} his
} hands in 96 hours.He lost his job. Later, after finding that message
} again,
} He mailed twenty copies. A few days later he got a better job.
}
} Dalan Fairchild received the message and, not believing - Threw the
} message away. Nine days later he died. In 1987, The message received by
} a young woman in California was very faded and barely readable. She
} promised herself that she would retype the message and send it on, But
} she set it aside to do it later. She was plagued with various problems,
} including expensive car repairs. The letter did not leave her hands
} within
} 96 hours. She finally typed the letter as promised and got a new car.
}
} Good Luck but please remember: 20 copies of this message







From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Thu, 1 Jun 1995 22:39:30 -0500 (CDT)
Subject: Ignore the Junk Mailer Please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi folks,..

I know it is an irratiation, but please ignore
the junk mailer. I have independently contacted
AOL and asked that they do something.

For the time being DONOT send him back any return
junk mail, It only aggrevates the problem.

Thanks.... Nestor




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Fri, 2 Jun 1995 5:29:59 -0500 (CDT)
Subject: Listserver Status....

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Fri, 2 Jun 1995 5:29:18 -0500 (CDT)
Subject: Listserver Status

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Colleagues

In order to reduce the junk mail you are getting
and we get the nut off-line, I have switched the
listerserver into manual mode. This means I read
and then forward all messages. Everything will
still get through it will only take awhile longer
since I must physically screen each posting.

You should not change your methodology. Continue
to post to:

Microscopy-at-AAEM.AMC.ANL.GOV

Nestor




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Fri, 2 Jun 1995 5:34:56 -0500 (CDT)
Subject: Re: Wrinkling acrylic resin.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Thu, 1 Jun 1995 23:01:19 -0700 (PDT)
Subject: Re: Wrinkling acrylic resin..

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer05.u.washington.edu

How were the slides cleaned? I've had problems that were traced back to
how well the slides were cleaned, even though the same subbing agent was
used. Chromic acid cleaned slides, subbed with gelatin are the only
thing that gets my Spurr's to lie flat when comverslipping with DPX. 50%
bleach, soaking slides 30 min., works almost as well, and also seems good
enough for AR (autoradiography, not the 'net definition). I put
methacrylate sections on bleach, chromic acid or slides cleaned with 1%
HCl in 95% ethanol, usually unsubbed, but also on chrome alum-gelatin or
pll subed slided.

Glen MacDonald
Hearing Development Laboratories
University of Washington
Seattle, WA 98195-3515
(206)543-8360
glenmac-at-u.washington.edu



On Thu, 1 Jun 1995 Goldmarker-at-aol.com wrote:

} A colleague is having difficulty preventing wrinkling of acrylic resin (2-3u
} thick) on glass slides. It does not seem to matter whether she uses
} pll-coated, gelatin-coated, or uncoated slides.
}
} She was told to add ethylene glycol to her formulation, but I am concerned
} about modifying the vendor's formulation.
}
} Does anyone have any suggestions?
}
} Thanks.
} Donald P. Cox, Ph.D. [goldmarker-at-aol.com]
}




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sun, 4 Jun 1995 10:10:18 -0500 (CDT)
Subject: RCA Photomultipliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: Petr Schauer :      petr-at-ISIBrno.Cz
Date: Fri, 2 Jun 1995 16:13:40 +0200
Subject: RCA Photomultipliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Networkers,

I am looking for a contact address (location, e-mail, fax, .....) of the
RCA company (photomultiplier tube division). Any help will be appreciated.

Petr
+-------------------------------------------+-------------------------+
| Dr. Petr Schauer | tel.: (+42 5) 41321246 |
| Head of Electron Microscopy Laboratory | fax : (+42 5) 746664 |
| ACADEMY OF SCIENCES OF THE CZECH REPUBLIC | (+42 5) 41211168 |
| INSTITUTE OF SCIENTIFIC INSTRUMENTS | E-mail: petr-at-isibrno.cz |
| Kralovopolska 147, CZ-612 64 Brno | petr-at-vabo.cz |
| Czech Republic | |
+-------------------------------------------+-------------------------+




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sun, 4 Jun 1995 10:11:06 -0500 (CDT)
Subject: XIVth Intnl Pfefferkorn Conference - update.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Information on this meeting is also available at:
http://www.engin.umich.edu/~jfmjfm/mas_folder/XIVPfefferkorn.html

John Mansfield
North Campus Electron Microbeam Analysis Laboratory
413 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX (313)936-3352
jfmjfm-at-engin.umich.edu, John.F.Mansfield-at-umich.edu
or jfmjfm-at-umich.edu they all reach me!
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html






From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sun, 4 Jun 1995 10:12:31 -0500 (CDT)
Subject: Micromaze Foreline Traps

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sun, 4 Jun 1995 10:13:03 -0500 (CDT)
Subject: Advice re: 35mm slidemaker

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: DCROMEY-at-CCIT.ARIZONA.EDU
Date: Fri, 02 Jun 1995 15:54:31 -0700 (MST)
Subject: Advice re: 35mm slidemaker

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are trying to decide between a Polaroid HR-6000 and a LaserGraphics
LFR-X 35mm slidemaker. Does anyone out there have any feedback on which
is better (and why) and if you have experience with either unit, what
kind of things do you like/dislike about it?

Thanks for your help.
Doug

.....................................................................
. Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy .
. Sr. Research Specialist University of Arizona .
. (office: LSN 451) 1501 N. Campbell Ave. .
. (voice: 520-626-2824) Tucson, AZ 85724 USA .
. (FAX: 520-626-2097) (email: dcromey-at-ccit.arizona.edu) .
.....................................................................





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sun, 4 Jun 1995 10:12:49 -0500 (CDT)
Subject: Micromaze Foreline Traps

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: dlb-at-aruba.ccit.arizona.edu (David Bentley)
Date: Fri, 2 Jun 1995 14:14:05 -0700
Subject: Micromaze foreline traps

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The manfacturers suggested regeneration cycle is correct if I
understand your trap correctly. I am guessing that the trap is filled with
molecular sieve (zeolite) which adsorbs oil and water vapors when cool.
During the heating phase, these vapors are driven off. Without a valve in
the line there is a good chance that the oil vapor will migrate towards the
diffusion pump (i.e. into the microscope). A valve in the line will avoid
this and any vapor will be expelled by the rotary pump. After a period of
time the molecular seive will lose efficiency, which heating will regenerate
its adsorption capacity. It may also be benificial to gas ballast the
rotary pump during regeneration. This will insure that water vapor which
will also be driven off during heating, does not contaminate the rotary pump
oil as much.

A concise answer to your questions
Yes, place a valve on the line
Heat on a weekly basis
It won't trap if its hot all the time
I would assume that the manufacturer either:
Put a thermostat in with the heating element
-OR-
Has a time versus temperature curve that is used


David Bentley
later
dlb





From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Sun, 4 Jun 1995 10:11:28 -0500 (CDT)
Subject: Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: meh-at-aretha.jax.org (Margaret Hogan)
Date: Fri, 02 Jun 1995 12:07:32 -0400
Subject: Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506021610.MAA02376-at-aretha.jax.org}
X-Sender: meh-at-aretha.jax.org
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


Histology - Biomedical Technologist/ Senior Biomedical Technologist

A regular full time position is open in The Jackson Laboratory Biological
Imaging Department - Histology Laboratory. The Jackson Laboratory is a
non-profit independent laboratory founded in 1929 on the premise that the
causes of cancer and other diseases could be discovered through
Mammalian genetic research. The Laboratory specializes in mammalian
genetics using inbred laboratory mice as model systems to study health
problems such as cancer diabetes, anemia, heart disease and aging.
Located on a large island in the gulf of Maine and surrounded by Acadia
National Park, The Jackson Laboratory is currently undergoing a major
expansion of its scientific staff and its research facilities.

Applicants with a Bachelors degree and two years related
laboratory experience working with Murine specimens preferred. The
position includes routine histological techniques ie paraffin embedding,
single and serial sectioning and cryotomy in conjunction with
Immunohistochemistry techniques, staining using heavy metals as well as
other special stains.

The successful candidate must be a self starter, pay attention to detail
and be able to work independently with little supervision.
This individual will be responsible for providing services in support of
numerous diverse research projects, must interact well with multiple
users and work productively in a team environment. The position includes
opportunities for advancement.

Salary range is mid to high $20,000 plus benefits and is negotiable
depending on level of experience.

Interested applicants should send CV to:

Joanne Bradt
Employment Specialist
The Jackson Laboratory
600 Main Street
Bar Harbor Maine 04609
(207) 288-3371 ext. 1281
(207) 288-3371 ext. 1082 FAX
jcb-at-aretha.jax.org





From: DCROMEY-at-CCIT.ARIZONA.EDU
Date: Fri, 02 Jun 1995 15:54:31 -0700 (MST)
Subject: Advice re: 35mm slidemaker

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy-at-AAEM.AMC.ANL.GOV, ZALUZEC-at-AAEM.AMC.ANL.GOV

Reply to: RE} Advice re: 35mm slidemaker
We have a Lazorgraphics LSR. The slides ore fine. However we
have had a problem with the shutter on the camera. One was
user or should I say miss-user problem. The second time the
shutter stoped opening in the middle of a role. It cost about $300
to get lazorgraphic to fix it the first time. (I am not sure how much a
new camera back would cost.) The second time was with in 90 days
of the first repair. Other than that any problems have been software
problem on our Mac or PC.
larry hawkey
hawkey-at-neuro.duke.edu

--------------------------------------


We are trying to decide between a Polaroid HR-6000 and a LaserGraphics
LFR-X 35mm slidemaker. Does anyone out there have any feedback on which
is better (and why) and if you have experience with either unit, what
kind of things do you like/dislike about it?

Thanks for your help.
Doug

.....................................................................
. Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy .
. Sr. Research Specialist University of Arizona .
. (office: LSN 451) 1501 N. Campbell Ave. .
. (voice: 520-626-2824) Tucson, AZ 85724 USA .
. (FAX: 520-626-2097) (email: dcromey-at-ccit.arizona.edu) .
.....................................................................








From: DCROMEY-at-CCIT.ARIZONA.EDU
Date: Fri, 02 Jun 1995 15:54:31 -0700 (MST)
Subject: Advice re: 35mm slidemaker

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506051223.IAA14808-at-neuro.duke.edu}
Microscopy-at-AAEM.AMC.ANL.GOV, ZALUZEC-at-AAEM.AMC.ANL.GOV

Reply to: RE} Advice re: 35mm slidemaker
We have a Lazorgraphics LSR. The slides ore fine. However we
have had a problem with the shutter on the camera. One was
user or should I say miss-user problem. The second time the
shutter stoped opening in the middle of a role. It cost about $300
to get lazorgraphic to fix it the first time. (I am not sure how much a
new camera back would cost.) The second time was with in 90 days
of the first repair. Other than that any problems have been software
problem on our Mac or PC.
larry hawkey
hawkey-at-neuro.duke.edu

--------------------------------------


We are trying to decide between a Polaroid HR-6000 and a LaserGraphics
LFR-X 35mm slidemaker. Does anyone out there have any feedback on which
is better (and why) and if you have experience with either unit, what
kind of things do you like/dislike about it?

Thanks for your help.
Doug

.....................................................................
. Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy .
. Sr. Research Specialist University of Arizona .
. (office: LSN 451) 1501 N. Campbell Ave. .
. (voice: 520-626-2824) Tucson, AZ 85724 USA .
. (FAX: 520-626-2097) (email: dcromey-at-ccit.arizona.edu) .
.....................................................................








From: A. Kent Christensen :      akc-at-umich.edu
Date: Mon, 5 Jun 1995 09:58:14 -0400 (EDT)
Subject: Re: Methyl cellulose solution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jeanne -- You can get this information in G. Griffith's "Fine Structure
Immunocytrochemistry", 1993, Springer-Verlag (Berlin, New York), ISBN
0-387-54805-X, pp. 175-177. -- A. Kent Christensen, Univ. of Michigan,
{akc-at-umich.edu} .

-----------------------------------

On 1 Jun 1995, Jeanne Barker wrote:

} REGARDING Methyl cellulose solution
} Hi,
} What is the best way to make a 2% aqueous solution of methyl cellulose?
} Should the water be cold, hot or RT? Does it require overnight spinning?
} Thanks in advances for your responses.
} Jeanne
}
}
}




From: DDKJoe-at-aol.com
Date: Mon, 5 Jun 1995 13:20:18 -0400
Subject: Re: Plastoid N..a source!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Don,
Plastoid N has had on-again, off-again availability from the manufacturer in
Germany. At the moment, we have an inventory and a source. We would be
happy to help your colleague.

Joe Tabeling
Delaware Diamond Knives
3825 Lancaster Pike
Wilmington, DE 19805
800-222-5143




From: _ _ _ _ _ MARK W. LUND _ _ _ _ _ :      LUNDM-at-PHYSC2.BYU.EDU
Date: Mon, 5 Jun 1995 11:58 MDT
Subject: Re: RCA Photomultipliers.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The RCA photomultplier company is now Burle Industries in
Lancaster PA. Tel: 717-296-6031
FAX 717-295-6096
regards
mark
---
Mark W. Lund, PhD
Director
MOXTEK, Inc.
Orem UT 84057




From: Strucural Biology Unit :      MICROSCOPY-at-SBSNOV1.AUCKLAND.AC.NZ
Date: Tue, 6 Jun 1995 17:16:24 GMT+1200
Subject: SEM versus TEM usage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Two years ago our zoology, botany, cell biology and biochemistry
departments formed a school of biological sciences (SBS). It made
little difference to the EM unit as it already serviced the TEM needs
of those departments. SEM on the other hand is located in
engineering. We are considering the purchase of an SEM as the
engineering Philips 505 is becomming very limiting as it is set up
primarily for engineering applications.

Our problem is estimating likely demand for a new (probably
environmental) SEM. We are currently running at about
5% SEM to 95% TEM. We suspect this is due to the SEM being unsuitable
for many biological applications, and its position 500 metres up the
road. We would like feedback from other Schools of Biological
science with both SEM and TEM as to the proportion of use each gets.

Ken and Terry


Microscopy Unit
School Of Biological Sciences
The University Of Auckland
Level 1, Thomas Building
Private Bag 92019
Auckland, NEW ZEALAND

phone:(09) 373 7999 ext 5986
fax: (09) 373 7417




From: Keith Moulding :      MCMOULDK-at-usthk.ust.hk
Date: 06 Jun 1995 17:39:01 +0800
Subject: Software - MaComis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Does anyone know where MaComis can be obtained from? The
program simulates dislocation contrast.

Thanks in advance.

Keith Moulding.

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Dr. Keith Moulding, ~
Materials Characterisation and Preparation Centre, ~
Hong Kong University of Science and Technology, ~
Clear Water Bay, ~
Kowloon, ~
Hong Kong. ~
~
Tel: (852) 2358 8724 ~
Fax: (852) 2358 2451 ~
~
E-mail: mcmouldk-at-usthk.ust.hk ~
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: {tealbe-at-ml.wpafb.af.mil}:ddn:wpafb
Date: 6-5-95 3:52pm
Subject: COMPUTER VIRUS ALERT

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential
Tribology ListServer {tribology-at-aaem.amc.anl.gov}
Message-Id: {950606082343.567-at-cliff.ml.wpafb.af.mil.0}
To: admin:ml:wpafb
Subj: COMPUTER VIRUS ALERT
Orig-Author: {Billy E. Teal {tealbe-at-ml.wpafb.af.mil} }:ddn:wpafb
-----------------------------------------------------------
COMPUTER VIRUS ALERT

Attention all DOS/Windows users. Someone is distributing
a file called PKZ300B.EXE or possibly PKZ300B.ZIP. This
is NOT a version of PKZIP, the most recent version is 2.04G.
These files are trojan horses and any attempt to execute them
will erase the C: drive. Please tell all your friends and
favorite BBS (Bulliten Board Services) about this hack.

Bill Teal


----- End of forwarded message -----




From: hmeekes-at-biosci.mbp.missouri.edu (Herman Meekes)
Date: Tue, 6 Jun 1995 14:05:11 -0500
Subject: Re: Software - MaComis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe microscopy {hmeekes-at-biosci.mbp.missouri.edu}






From: emlab-at-ucsco.ucsc.edu (Jon Krupp)
Date: Tue, 6 Jun 1995 16:21:06 -0700
Subject: Help Identifying Nikon Scope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506062320.QAA02795-at-ucsco.ucsc.edu}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I need some help identifying a Nikon compound microscope. It belongs to a
retired professor who used it for a few years and now needs to know how to
describe it for possible sale. Of course the original documentation has
been misplaced and he does not remember the details of the original order.
So, in addition to trying to get help from Nikon, I thought I would try
here too.

The instrument was purchased in 1984 for about $12,000. It looks like a
standard compound microscope on a larger than normal stand. The base is
about 18" square and it looks like the microscope body can be focused up
about 6" to 10" from the stage. It has both transmitted and reeflected
light sources provided by fiber optics.

The stage is this instruments claim to fame according to its owner. It is
large, big enough to hold an oversize thin section of rock maybe 6" x 6".
It has precision XY controls that somehow interacted with a computer
through a Nikon Digital Counter to record coordinates. The stage drives are
manual, but look like they could enclose motor drives.

That's about all he can remember, if you think you can help, please let me know.

Thanks

Jonathan Krupp
Microscopy and Imaging Lab
University of California
Santa Cruz, CA 95064
(408) 459-2477
emlab-at-ucsco.ucsc.edu






From: :      ARRELL-at-jrc.nl
Date: Wed, 7 Jun 1995 10:37:27 GMT+0200
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.950607103727.608-at-FS-IAM-1.JRC.NL}

UNSUBSCRIBE ARRELL-at-JRC.NL




From: dietmar.reiter-at-uibk.ac.at (Dietmar Reiter)
Date: Wed, 7 Jun 1995 12:58:26 +0100
Subject: water soluble embedding media

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: c71956-at-u1.uibk.ac.at
Message-Id: {v01510100abfb436e4d89-at-[138.232.71.39]}
Mime-Version: 1.0
Content-Type: text/plain; charset=us-ascii

For embedding of small invertebrates in whole-mount immunocytochestry:
which water soluble embedding media - preserving fluorescence is not
necessary - are recommended?

Thanks for any suggestions, -Dietmar-


+++ Dietmar Reiter, Dept. of Zoology and Limnology
+++ University of Innsbruck
+++ Technikerstrasse 25, A-6020 Innsbruck, Austria
+++ phone: (43)-512-507 ext. -6170, fax ext. -2930






From: :      ARRELL-at-jrc.nl
Date: Wed, 7 Jun 1995 13:21:14 GMT+0200
Subject: UNSUBSCRIBE ARRELL@JRC.NL

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.950607132114.256-at-FS-IAM-1.JRC.NL}

UNSUBSCRIBE ARRELL-at-JRC.NL




From: Mev H van der Merwe :      HVDM-at-op1.up.ac.za
Date: Wed, 7 Jun 1995 15:51:54 GMT+2
Subject: staining problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hallo

I am using the Immuno-Bed kit for Histology. When staining the
Histology section, I get a coloured back ground in the plastic. Can
someone please help me?

Thanks

E.mail: HvdM-at-opl.up.ac.za




From: Alec Madsen :      alecm-at-u.washington.edu
Date: Wed, 7 Jun 1995 08:49:23 -0700 (PDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



UNSUBSCRIBE ALECM-at-U.WASHINGTON.EDU




From: hukee.margaret-at-mayo.EDU (Marge Hukee)
Date: Wed, 7 Jun 1995 11:07:52 +0200
Subject: listserver vs microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506071606.AA20134-at-fermat.Mayo.EDU}
X-Sender: hukem-at-fermat.mayo.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear subscribers;

I grow weary of subscribe and unsubscribe messages. Please send these
commands to listserver-at-aaem.amc.anl.gov and NOT to
microscopy-at-aaem.amc.anl.gov

Thanks
This message is small enough to post by your machine

Marge Hukee
EM Core Facility hukee.margaret-at-mayo.edu
Mayo Foundation (507) 284-3148
----------------------------------------------------------------------------
--






From: Elinor Solit :      cambrex-at-world.std.com
Date: Wed, 7 Jun 1995 12:19:25 +0059 (EDT)
Subject: Re: Help Identifying Nikon Scope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jon,

I suggest your friend contact Nikon in Long Island, New York. Their
number is 516-547-8500. Ask for any of their Product Managers, they are
all capable.

Hope this helps.

Ellie Solit
The Microscope Book

On Tue, 6 Jun 1995, Jon Krupp wrote:

} I need some help identifying a Nikon compound microscope. It belongs to a
} retired professor who used it for a few years and now needs to know how to
} describe it for possible sale. Of course the original documentation has
} been misplaced and he does not remember the details of the original order.
} So, in addition to trying to get help from Nikon, I thought I would try
} here too.
}
} The instrument was purchased in 1984 for about $12,000. It looks like a
} standard compound microscope on a larger than normal stand. The base is
} about 18" square and it looks like the microscope body can be focused up
} about 6" to 10" from the stage. It has both transmitted and reeflected
} light sources provided by fiber optics.
}
} The stage is this instruments claim to fame according to its owner. It is
} large, big enough to hold an oversize thin section of rock maybe 6" x 6".
} It has precision XY controls that somehow interacted with a computer
} through a Nikon Digital Counter to record coordinates. The stage drives are
} manual, but look like they could enclose motor drives.
}
} That's about all he can remember, if you think you can help, please let me know.
}
} Thanks
}
} Jonathan Krupp
} Microscopy and Imaging Lab
} University of California
} Santa Cruz, CA 95064
} (408) 459-2477
} emlab-at-ucsco.ucsc.edu
}
}




From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Wed, 7 Jun 1995 11:50:23 -0700 (PDT)
Subject: Re: staining problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer01.u.washington.edu

What stain are you using? I often destain with 35% ethanol to wash off
the stain, followed by distilled water do eliminate spotting left behind
by the ethanol.


Glen MacDonald
Hearing Development Laboratories
University of Washington
Seattle, WA 98195-3515
(206)543-8360
glenmac-at-u.washington.edu


On Wed, 7 Jun 1995, Mev H van der Merwe wrote:

} Hallo
}
} I am using the Immuno-Bed kit for Histology. When staining the
} Histology section, I get a coloured back ground in the plastic. Can
} someone please help me?
}
} Thanks
}
} E.mail: HvdM-at-opl.up.ac.za
}




From: Fred Hayes :      fahayes-at-ucdavis.edu
Date: Wed, 7 Jun 1995 15:53:38 -0700 (PDT)
Subject: sectioning polyester filters for LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Does anyone have experience thick sectioning the "Transwell Clear"
filters from Corning Costar? They thin section (80nm) fine without any
problems, but appear to have no elasticity at the light level(1000nm).
This is a 12mm polyester filter with cultured RPE cells. The customer
wants a cross sectional view since he is interested in quantifying
basement membrane deposition. The filters are embedded in EMbed 812 with
plenty of resin on both sides of the filter. Block face measures 0.5 x 5.0mm
Blocks or filter discs are bisected with that edge being sectioned.

1) the problem is section wrinkling since the filter refuses to stretch
with the resin and monolayer. I've tried differing the temp, using acetone,
toulene, Etoh, and Chloroform. Is there a solvent that permeates or
softens polyester?

Fred A. Hayes 916-752-7712 work
University of California,Davis 916-752-4701 work
School of Medicine
Department ofMedical Pathology; EM Lab
MSIA E-mail:
Davis, CA 95616 fahayes-at-ucdavis.edu

1320 Dogwood Court 916-678-6280 home
Dixon, CA 95620-3227







From: m.dickson-at-unsw.edu.au (Melvyn R. Dickson)
Date: Thu, 8 Jun 1995 10:27:41
Subject: Re: SEM-x.s. microporous PTFE membrane

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-aaem.amc.anl.gov



} Hello:

} First use of the forum... Is there is a technique for preparing cross
} section profiles of microporous, multi-layered PTFE membranes
} approximately 10 to 50 micrometers thick for FE-SEM.

We do a lot of microporous membranes for FESEM examination. Mostly they crack
quite well if we chill them under liquid nitrogen and crack them with pre
cooled forceps. But I have also had to tear them and at times chill them and
chop them with a chilled razor blade hit with a hammer. We have a Balzers BAF
400 but we seldom use it for cross fractures. After fracturing we use around
2 nm of chromium to coat.

For studying penetration down into pores we just cross sectioned epoxy
embedded membranes and looked at them with TEM. Its the only way if membranes
are still flexible at Liquid N2 temperatures. Anyway I distrust procedures
that remove material as I reckon the fouling material / coating material
whatever is likely to move also.

Mel Dickson




From: Mev H van der Merwe :      HVDM-at-op1.up.ac.za
Date: Thu, 8 Jun 1995 13:13:05 GMT+2
Subject: Glass knives for Immuno-Bed sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I must cut Immuno-Bed sections 25mm x 20mm. I have a problem with my
steel knife. I always get scars on the sections. I sharpend my
knife at an angle of 36 and for the finishing process 5-10 minutes
at an angle of 35. I sharpend the knife for 4-5 days. I have glass
for a ralph knife, but this glass is too soft and only suitable for
wax sections. The surface of E/M glass is too small, I prefer
cutting with a glass knife. Can someone please help me?

Thanks
Hildagoenda v.d. Merwe


E-MAIL: HvdM-at-opl.ac.za




From: Mev H van der Merwe :      HVDM-at-op1.up.ac.za
Date: Thu, 8 Jun 1995 14:45:24 GMT+2
Subject: glass knives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I must cut Immuno-Bed sections 25mm x 20mm. I have a problem with my
steel knife. I always get scars on the sections. I sharpend my
knife at an angle of 36 and for the finishing process 5-10 minutes at
an angle of 35. I sharpend the knife for 4-5 days.

I have glass for a ralph knife but this glass is too soft and only
suitable for wax sections. The surface of E/M glass is too small.

I prefer cutting with a glass knife. Can someone help please.

Thanks

Hildagoenda v.d. Merwe
University of Pretoria
South Africa


E-mail: HVDM-at-opl.up.ac.za




From: MacisNo1-at-aol.com
Date: Thu, 8 Jun 1995 09:30:50 -0400
Subject: Re: Help with Nikon Scope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jonathan Krupp,

It sounds to me like either UM or TM20. These were large pillar type
"Toolmakers" microscopes or also known as Measuring scopes due to the long
throat (to accommodate large specimens) and accurate linear stages. Some were
manual as yours with micrometers and others use scales with LED readouts. In
any event. let me know what color the stand is and that will help me further
identify it.

Sincerely,

Larry Kordon
Nikon, Inc.
MacisNo1-at-aol.com




From: farmer-at-emlab.zoo.uga.edu
Date: Thu, 8 Jun 1995 10:11:50 +0000
Subject: Glycogen in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506081411.AA14319-at-emlab.zoo.uga.edu}
Comments: Authenticated sender is {farmer-at-emlab.zoo.uga.edu}

Does anyone have a good method for positively staining glycogen in
the TEM? I am aware of glycogen's staining characteristics when it
reacts with lead salts and the importance of OsO4 pH in affecting the
appearance of glycogen. Unfortunately these and other techniques
(e.g. agglutination with Con A) are not specific enough for this
project. We need to demonstrate glycogen and nothing but glycogen.

Are there any commercially available enzyme-colloidal gold complexes
that might be of use? Any references or advice would be
appreciated.

Mark Farmer

Mark A. Farmer
Director, Ctr. Ultrastructural Research
University of Georgia, Athens, GA 30602
(706)542-4080 Voice (706)542-4271 FAX
farmer-at-emlab.zoo.uga.edu





From: Paul Webster :      Paul.Webster-at-quickmail.cis.yale.edu
Date: 8 Jun 1995 11:00:58 -0400
Subject: Glycogen in TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1409513192.93718-at-QuickMail.Yale.edu}

Use antibodies to glycogen.





From: hukee.margaret-at-mayo.EDU (Marge Hukee)
Date: Thu, 8 Jun 1995 10:52:38 +0200
Subject: sectioning immuno-bed (25mmX20mm)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506081550.AA00995-at-fermat.Mayo.EDU}
X-Sender: hukem-at-fermat.mayo.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We used to section JB-4 with the fresh broken edge of a microscope slide
that was inserted into a razor blade adapter. The older ultratomes had
these adapters, but with the new machines you may have to have a machinist
make one for you. A machine that uses metal knives usually includes a
holder for the disposable knives that may be also adaptable.
For what it's worth
Marge

Marge Hukee
EM Core Facility hukee.margaret-at-mayo.edu
Mayo Foundation (507) 284-3148
----------------------------------------------------------------------------
--






From: masur-at-msvax.mssm.edu
Date: Thu, 08 Jun 1995 13:56:41 -0500
Subject: double stain with two mouse monoclonals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Has anyone had success with immunofluorescent co-localization of two
antigens, on one coverslip of cultured cells using using two different
mouse monoclonal antibodies. (we have had results in which we could not be
certain that we had successfully blocked after the first antibody)

protocol please

Sandra K. Masur Box 1183
fax: 212-289-5945 Mount Sinai School of Medicine
phone: 212-241-6544 or 0089 NY NY 10029-6574
e-mail:masur-at-msvax.mssm.edu









From: Paul Webster :      Paul.Webster-at-quickmail.cis.yale.edu
Date: 8 Jun 1995 15:00:07 -0400
Subject: Double stain with two mouse

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1409498732.63153-at-QuickMail.Yale.edu}

Sandra K. Masur writes:
Has anyone had success with immunofluorescent co-localization of two
antigens, on one coverslip of cultured cells using using two different
mouse monoclonal antibodies. (we have had results in which we could not be
certain that we had successfully blocked after the first antibody).

It would be interesting to know how you blocked after the first antibody. There
is no blocking protocol that I know of which will stop antibodies binding to
antigens.

As far as I know there is no published protocol for double labeling with two
monoclonal antibodies on the same preparation. The normal way is to directly
label the antibodies with fluorescent tags and not use secondary antibodies. It
should also be easy to label one of the monoclonals with a small antigenic
molecule such as biotin and detect it with anti-biotin antibodies (or
FITC-streptavidin).

If you were using sections I would suggest cutting serial sections, labeling one
set with the first antibody and the other with the second.








From: Michael Rock :      merock-at-u.washington.edu
Date: Thu, 8 Jun 1995 09:32:09 -0700 (PDT)
Subject: Re: sectioning polyester filters for LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fred
you might try using a vacuum dessicator to remove the wrinkles, it seem
to work well for 1 micron sections on glass, it may work for TEM as well.
-Mike

On Wed, 7 Jun 1995, Fred Hayes wrote:

}
} Does anyone have experience thick sectioning the "Transwell Clear"
} filters from Corning Costar? They thin section (80nm) fine without any
} problems, but appear to have no elasticity at the light level(1000nm).
} This is a 12mm polyester filter with cultured RPE cells. The customer
} wants a cross sectional view since he is interested in quantifying
} basement membrane deposition. The filters are embedded in EMbed 812 with
} plenty of resin on both sides of the filter. Block face measures 0.5 x 5.0mm
} Blocks or filter discs are bisected with that edge being sectioned.
}
} 1) the problem is section wrinkling since the filter refuses to stretch
} with the resin and monolayer. I've tried differing the temp, using acetone,
} toulene, Etoh, and Chloroform. Is there a solvent that permeates or
} softens polyester?
}
} Fred A. Hayes 916-752-7712 work
} University of California,Davis 916-752-4701 work
} School of Medicine
} Department ofMedical Pathology; EM Lab
} MSIA E-mail:
} Davis, CA 95616 fahayes-at-ucdavis.edu
}
} 1320 Dogwood Court 916-678-6280 home
} Dixon, CA 95620-3227
}
}
}
}




From: hukee.margaret-at-mayo.EDU (Marge Hukee)
Date: Thu, 8 Jun 1995 16:36:16 +0200
Subject: DOUBLE STAIN

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fixatives generally will alter the ability of an antibody to react to a
second antibody. This can be easily tested by trying to label your first
antigen after pretreatment with fixative. A paper is available in the
literature describing a blockage of first antibody with vapor from
formaldehyde fixative generated at 80 degrees C and subsequent labeling
with a second antibody of the same specie. This paper is by Wang and
Larsson and can be found in Histochemistry bol.83:page 47, 1985. I have
used this method successfully at the electron microscopy level. If I can
be of any further help, please feel free to contact me.
Marge

Marge Hukee
EM Core Facility hukee.margaret-at-mayo.edu
Mayo Foundation (507) 284-3148
----------------------------------------------------------------------------
--






From: David Garrett :      DGARRETT-at-gab.unt.edu
Date: Thu, 8 Jun 1995 16:46:15 CST6CDT
Subject: Question: Re: sectioning polyester filters for LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Date sent: Thu, 8 Jun 1995 09:32:09 -0700 (PDT)
} From: Michael Rock {merock-at-u.washington.edu}
} To: Fred Hayes {fahayes-at-ucdavis.edu}
} Copies to: microscopy-at-aaem.amc.anl.gov
} Subject: Re: sectioning polyester filters for LM

} Fred
} you might try using a vacuum dessicator to remove the wrinkles, it seem
} to work well for 1 micron sections on glass, it may work for TEM as well.
} -Mike
}
Please explain.
Thanks
***************************************
David Garrett "DGARRETT-at-GAB.UNT.EDU"
University of North Texas
Dept. Biological Sciences
(817)565-3964 Fax (817)565-4136
***************************************




From: hukee.margaret-at-mayo.EDU (Marge Hukee)
Date: Thu, 8 Jun 1995 17:19:03 +0200
Subject: glycogen staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506082217.AA10481-at-fermat.Mayo.EDU}
X-Sender: hukem-at-fermat.mayo.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear Mark,
Glycogen can be stained by adding potassium ferricyanide to primary and
secondary fixatives. If you wish to differentiate glycogen B-particles
from ribosomes, do not use lead stain on grids. References are available
in the literature. Two that I know of are: Duvall, Hukee in Ann. Otol.
Rhinol. Laryngol. vol 85:234, 1976 this technique was modified from De
Bruijn in Histochem Journal vol. 8:121,1976 or an earlier paper by this
same author. In more recent times De Bruijn has looked at various
fixatives with x-ray microanalysis in Histochemical Journal vol.13:125,
1981.

Marge Hukee
EM Core Facility hukee.margaret-at-mayo.edu
Mayo Foundation (507) 284-3148
----------------------------------------------------------------------------
--






From: images1-at-biosci.mbp.missouri.edu (Michael Stanley)
Date: Thu, 8 Jun 1995 05:26:22 -0500
Subject: Double stain with two mouse

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510101abfc7e3931ae-at-[128.206.15.185]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Date: 8 Jun 1995 15:00:07 -0400
} From: "Paul Webster" {Paul.Webster-at-quickmail.cis.yale.edu}
} Subject: Double stain with two mouse
} To: "Microscopy Bulletin Board" {microscopy-at-AAEM.AMC.ANL.GOV}
}
} Sandra K. Masur writes:
} Has anyone had success with immunofluorescent co-localization of two
} antigens, on one coverslip of cultured cells using using two different
} mouse monoclonal antibodies. (we have had results in which we could not be
} certain that we had successfully blocked after the first antibody).
}
} It would be interesting to know how you blocked after the first antibody.
} There
} is no blocking protocol that I know of which will stop antibodies binding to
} antigens.
}
} As far as I know there is no published protocol for double labeling with two
} monoclonal antibodies on the same preparation. The normal way is to directly
} label the antibodies with fluorescent tags and not use secondary antibodies. It
} should also be easy to label one of the monoclonals with a small antigenic
} molecule such as biotin and detect it with anti-biotin antibodies (or
} FITC-streptavidin).
}
} If you were using sections I would suggest cutting serial sections,
} labeling one
} set with the first antibody and the other with the second.
}
}
There are a couple of papers available: Stefan Wurden and Uwe Homberg, The
Journal of Histochem and Cytochem, vol 41, pg 627, 1993. And S.A.L. Carl,
Gillete-Ferguson, and D.G. Fertuson, J of Histochem and Cytochem, vol 41,
pg 1273, 1993.

We have a user that finally managed to get a protocol to work but it took
months...

C. Michael Stanley, Ph.D.
Coordinator, Associate Director
Molecular Cytology Core Facility
Molecular Biology Program
2 Tucker Hall
University of Missouri-Columbia
Columbia, MO. 65211
(314) 882-4895
fax= 314-882-0123






From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Thu, 8 Jun 1995 16:36:08 -0500
Subject: Re: Double stain with two mouse

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There are, in fact, several articles in which a single section has been
stained with 2 monoclonals from the same species. e.g., Lewis-Carl et al.,
J. Histochem. Cytochem 1993 41:1273-1278
Boorsma Histochemistry 1984; 80:103
an der Loos e tal., 1987 J. Histochem Cytochem 1987 35:1139
Carl et al, J. Histochem Cytochem 1993 41:1273
and a bunch of others.
See these papers for the controls that have been used.
An alternative approach is to stain a section with a monoclonal,
photograph, strip the antigen and re-stain with a second antibody. We have
an abstract at the coming EMSA meeting on this approach (Stanley and
Phillips) but others have also tried this approach: one example is Lan et
al., J. Histochem. Cytochem 1995, 43:97
}
} } Sandra K. Masur writes:
} } Has anyone had success with immunofluorescent co-localization of two
} } antigens, on one coverslip of cultured cells using using two different
} } mouse monoclonal antibodies. (we have had results in which we could not be
} } certain that we had successfully blocked after the first antibody).
} }
} } It would be interesting to know how you blocked after the first antibody.
} } There
} } is no blocking protocol that I know of which will stop antibodies binding to
} } antigens.
} }
} } As far as I know there is no published protocol for double labeling with two
} } monoclonal antibodies on the same preparation. The normal way is to directly
} } label the antibodies with fluorescent tags and not use secondary
} } antibodies. It
} } should also be easy to label one of the monoclonals with a small antigenic
} } molecule such as biotin and detect it with anti-biotin antibodies (or
} } FITC-streptavidin).
} }
} } If you were using sections I would suggest cutting serial sections,
} } labeling one
} } set with the first antibody and the other with the second.
} }
}

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: lmiller-at-ux1.cso.uiuc.edu (Lou Ann Miller)
Date: Thu, 8 Jun 1995 20:59:53 -0700
Subject: Re: glycogen staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510101abfd76ff648a-at-[128.174.23.104]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi Mark,

Just a little note to this,

Be sure NOT to enbloc stain with Uranyl Acetate after the KCN if using UA
is your standard procedure. UA will pull the glycogen back out of your
sample.

Lou Ann



} Dear Mark,
} Glycogen can be stained by adding potassium ferricyanide to primary and
} secondary fixatives. If you wish to differentiate glycogen B-particles
} from ribosomes, do not use lead stain on grids. References are available
} in the literature. Two that I know of are: Duvall, Hukee in Ann. Otol.
} Rhinol. Laryngol. vol 85:234, 1976 this technique was modified from De
} Bruijn in Histochem Journal vol. 8:121,1976 or an earlier paper by this
} same author. In more recent times De Bruijn has looked at various
} fixatives with x-ray microanalysis in Histochemical Journal vol.13:125,
} 1981.
}
} Marge Hukee
} EM Core Facility hukee.margaret-at-mayo.edu
} Mayo Foundation (507) 284-3148
} ----------------------------------------------------------------------------
} --

***************************
Lou Ann Miller MT(ASCP) CT(MSA)
Microscopic Imaging Lab
College of Vet. Medicine
University of Illinois
2001 S Lincoln Ave
Urbana,Illinois 61801
217-244-1566
lmiller-at-ux1.cso.uiuc.edu






From: A. HONARBAKHSH-RAOUF :      CER5AH-at-ECU-01.NOVELL.LEEDS.AC.UK
Date: Fri, 9 Jun 1995 08:10:27 GMT
Subject: cast iron thin foil

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello
Does anyone know any suitable solution for preparation of thin
foil using jet polishing of ductile cast iron(presence of two
different phases ie free graphite and matrix)? Is there any other
method except ion beam?
Thanks
Abbas




From: Neuberger, Damian
Date: Thursday, June 08, 1995 4:18PM
Subject: MRS-3 Mag Standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Has anybody used the Geller MicroAnalytical MRS-3XYZ NIST
traceable magnification standard. It is reported to be traceable in XY
and Z axes and can be used for SEM and optical microscopes
(including confocal in reflected mode?). Comments on experience
with its use and performance would be much appreciated.

Damian Neuberger
neuberd-at-roundlake.baxter.com




From: Witold Zielinski :      WIZIEL-at-inmat.pw.edu.pl
Date: Fri, 9 Jun 1995 14:56:49 CET
Subject: Re: cast iron thin foil

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


*From: "A. HONARBAKHSH-RAOUF" {CER5AH-at-ECU-01.NOVELL.LEEDS.AC.UK}
*To: microscopy-at-aaem.amc.anl.gov
*Date sent: Fri, 9 Jun 1995 08:10:27 GMT
*Subject: cast iron thin foil
*Priority: normal

*Hello
*Does anyone know any suitable solution for preparation of thin
*foil using jet polishing of ductile cast iron(presence of two
*different phases ie free graphite and matrix)? Is there any other
*method except ion beam?
*Thanks
*Abbas

If the particles are small, like mikron size, try mixture of
95% acetic accid and 5% perchloric accid. Room temp! and voltage
about 80 - 90 Volts.
Good luck
Witold Zielinski
Warsaw University of Tech.
Warszawa




From: gbza40-at-udcf.gla.ac.uk
Date: Fri, 9 Jun 1995 14:14:44 +0100
Subject: MORE ON GLYCOGEN STAINS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9652.199506091314-at-lenzie.cent.gla.ac.uk}

Dear Mark Farmer

No one's mentioned the Thiery on-section method for carbohydrate staining
which shows up glycogen extremely well through silver deposition on
unstained epoxy sections. It may be an old method but it is very precise
and allows resolution of glycogen substructure if the reaction is time
varied to give optimal contrast. There's usually little background at
moderate magnifications and it works on routine glut/OsO4 blocks thus giving
good ultrastructure for this kind of electron cytochemistry.

Original paper is Thiery (1967), but the method is described in Hayat's book
Positive Staining for EM (1975) p111 and in many other texts also.

Whether this is sufficient for your needs is for you to decide !

Good luck

Laurence Tetley
EM Centre Biological Lab
Institute of Biomed and Life Sciences
University of Glasgow
Scotland, UK





From: Goldmarker-at-aol.com
Date: Fri, 9 Jun 1995 11:00:19 -0400
Subject: In-Situ hybridization w/ gold?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a good reference(s) and/or protocol for the use of
anti-digoxigenin gold or anti-biotin gold at either the EM or LM level?

I am trying to gather as much material as possible to help others with
in-situ hybridization with gold techniques.

In addition, if anyone would like to have a "direct" and "indirect" SAD-gold
(1nm) procedure, I'll be happy to send it.

Regards, Donald P. Cox, Ph.D.
Goldmark Biologicals
437 Lock St
Phillipsburg, NJ 08865
(908) 859-2631 - - - (908) 859-2875-FAX
E-mail: goldmarker-at-aol.com




From: David Henriks :      73531.1344-at-compuserve.com
Date: 09 Jun 95 11:29:59 EDT
Subject: cast iron thin foil

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You are certainly trying to thin some difficult material!

A good starting point would be:

65ml hydrochloric acid
435 ml methanol
20 ml butyl cellosolve

Temperature: -20 degrees C
Voltage: 110V
Current: 120 mA

These conditions are as used on a South Bay Technoilogy Model 550 Single
Vertical Jet Electropolisher. If you are using a twin-jet system, you will need
to adjust some of the parameters, but the basic solution is a good starting
point.

You may want to refere to the following paper:

"Technique for Jet Thinning Aged Iron-Chromium Alloys for TEM"
B.J. Kestel
Ultramicroscopy 25 (1988) 89-90

If you prefer, I could send you a copy of the paper at no charge. Of course, I
would also be pleased to send you information on our jet polisher and/or any of
our other sample preparation products.

Best regards-





From: David Henriks, 73531,1344
Date: 09 Jun 95 11:34:14 EDT
Subject: Copy of: cast iron thin foil

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



---------- Forwarded Message ----------


RE: Copy of: cast iron thin foil

You are certainly trying to thin some difficult material!

A good starting point would be:

65ml hydrochloric acid
435 ml methanol
20 ml butyl cellosolve

Temperature: -20 degrees C
Voltage: 110V
Current: 120 mA

These conditions are as used on a South Bay Technoilogy Model 550 Single
Vertical Jet Electropolisher. If you are using a twin-jet system, you will need
to adjust some of the parameters, but the basic solution is a good starting
point.

You may want to refere to the following paper:

"Technique for Jet Thinning Aged Iron-Chromium Alloys for TEM"
B.J. Kestel
Ultramicroscopy 25 (1988) 89-90

If you prefer, I could send you a copy of the paper at no charge. Of course, I
would also be pleased to send you information on our jet polisher and/or any of
our other sample preparation products.

Best regards-

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com





From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Fri, 9 Jun 1995 11:43:44 -0600
Subject: allEM: Geller Mag Standard MRS-3XYZ NIST

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Anyone have ordering information & pricing on the Geller MicroAnalytical
MRS-3XYZ NIST magnification standard? Thank you.


John J. Bozzola
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
Phone: 618-453-3730
Fax: -2665
Email: bozzola-at-siu.edu OR bozzola-at-qm.c-cem.siu.edu






From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Fri, 9 Jun 1995 14:41:38 -0600
Subject: TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Regarding hi res CCD cameras (1024x1024) for the TEM (Hitachi H7100), does
anyone know of a reliable vendor who might be able to supply & install a
working camera within 4-6 weeks? Thank you kindly.

John J. Bozzola
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
Phone: 618-453-3730
Fax: -2665
Email: bozzola-at-siu.edu OR bozzola-at-qm.c-cem.siu.edu






From: Eugene Bluhm :      bluhm-at-umr.edu
Date: Fri, 9 Jun 1995 15:07:37 -0600 (CDT)
Subject: subscribe bluhm@umr.edu

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe bluhm-at-umr.edu




From: bozzola
Date: Friday, June 09, 1995 11:43AM
Subject: allEM: Geller Mag Standard MRS-3XYZ NIST

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Anyone have ordering information & pricing on the Geller MicroAnalytical
MRS-3XYZ NIST magnification standard? Thank you.


John J. Bozzola
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
Phone: 618-453-3730
Fax: -2665
Email: bozzola-at-siu.edu OR bozzola-at-qm.c-cem.siu.edu






From: tvoiles-at-unlinfo.unl.edu (Todd Voiles)
Date: Fri, 09 Jun 1995 14:24:21 -0400
Subject: Getting Users to Use Equipment Properly

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506091923.AA28231-at-unlinfo.unl.edu}
X-Sender: tvoiles-at-129.93.1.11 (Unverified)
X-Mailer: Windows Eudora Version 1.4.4
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

He at U of Nebraska we have been running into the re-occuring problem
of users of the scopes and lab being overly messy and careless with
the equipment. We've tried posting signs about it, writing standard and
emergency operating proccedures to no avail. I'm sure other people
have had similar problems in their labs........anybody want to share
some solutions?


Todd Voiles
Facility Manager
Central Facility for Electron Microscopy
Center for Materials Research and Analysis

University of Nebraska at Lincoln

tvoiles-at-unlinfo.unl.edu





From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Fri, 9 Jun 1995 16:14:17 -0500
Subject: *M/fixatives/pfa

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
Anyone have a protocol for dissolving powder paraformaldehyde in
acetone???
Thanks,
Tobias Baskin

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
___ ____ ^ ____ _____ Tobias I. Baskin
/ \ / / \ / \ / University of Missouri
/ | / / \ / / Biological Sciences
/___ / /__ /_____\ / /__ 109 Tucker Hall
/ / / \ ( / Columbia, MO 65211 USA
/ / / \ \ / voice: 314-882-0173
/ /____ / \ \____/ /_____ fax: 314-882-0123






From: Joseph P. Neilly 708-938-5024 :      NEILLY.JOSEPH-at-igate.pprd.abbott.com
Date: Fri, 09 Jun 1995 08:14:00 -0600 (CST)
Subject: Re:MRS-3 Mag Standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta RANDB; Relayed; Fri, 09 Jun 1995 10:36:22 -0600
Mr-Received: by mta MCM$RAND; Relayed; Fri, 09 Jun 1995 10:36:24 -0600
Mr-Received: by mta RANDC; Relayed; Fri, 09 Jun 1995 10:36:37 -0600
Alternate-Recipient: prohibited
Disclose-Recipients: prohibited
Content-Return: prohibited

Damian,

We purchased the MRS-2 a couple of years ago and use it routinely for
both SEM and optical microscopy (both reflected and transmitted). I
believe Geller has added some video patterns on the MRS-3 but the basic
pattern has not changed. The best feature of this standard is the
ability to do X and Y checks with a single image. We do have a little
difficulty getting good contrast at low kV (2 kV) in the SEM in the
secondary mode. The only other problem we have with this standard is the
fact that it is so useful for both SEM and optical that it never stays
in the same lab very long.

Joe Neilly
Cellular and Microscopic Research
Abbott Laboratories
Abbott Park, IL
E-mail: neilly.joseph-at-igate.abbott.com
Phone: (708)-938-5024






From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Fri, 9 Jun 1995 18:39:11 -0500
Subject: Re: Getting Users to Use Equipment Properly

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In message {9506091923.AA28231-at-unlinfo.unl.edu} Todd Voiles writes:
} He at U of Nebraska we have been running into the re-occuring problem
} of users of the scopes and lab being overly messy and careless with
} the equipment. We've tried posting signs about it, writing standard and
} emergency operating proccedures to no avail. I'm sure other people
} have had similar problems in their labs........anybody want to share
} some solutions?



Its hopeless, Todd. I recognize your malady right off the bat. You've got dem
ol' EM Lab manager blues. You must learn to float above it all with equanimity,
for your own peace of mind, and keep your service contracts current. As one
hoary old sage of EM put it: "Be in the lab, but not of it". There's nothing
like managing a service lab to sharpen up your spiritual practice.

Stay cool!

Gib

--

Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu

"MICROSCOPY & MICROANALYSIS 96" in Minneapolis, Minnesota, Aug.11-15, 1996





From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Fri, 9 Jun 1995 19:01:58 -0500
Subject: Re: Getting Users to Use Equipment Properly

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In message {9506091923.AA28231-at-unlinfo.unl.edu} Todd Voiles writes:
} He at U of Nebraska we have been running into the re-occuring problem
} of users of the scopes and lab being overly messy and careless with
} the equipment. We've tried posting signs about it, writing standard and
} emergency operating proccedures to no avail. I'm sure other people
} have had similar problems in their labs........anybody want to share
} some solutions?

But seriously, Todd (tho the thoughts I sent earlier ARE part of my lab attitude
part of the time) you only mention posting signs and writing operating
procedures for your lab users. Thats fine and necessary, but it really takes
face to face discussion with users about whats expected, setting the tone, and
the written stuff is just to remind them of that when you are not around. Get
your lab director or department head or some other heavyweight authority figure
to back you up, make it clear to users that they may be asked to leave if they
cannot follow a few reasonable operating procedures. In my experience, most will
do that if approached in the right way, but there are always a few...........

Sometimes I get better results with younger people, students. Just screw off the
top of their head, pour in your information, screw their cover back on, give
them a pat on the butt and they are off and running. Sometimes its the older
experienced, rusty crusty heads whose lab habits are set that are the difficult
ones to communicate with. Take a workshop on management to get some ideas.

Good luck. Gib



--

Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu

"MICROSCOPY & MICROANALYSIS 96" in Minneapolis, Minnesota, Aug.11-15, 1996





From: AWBlackwoo-at-aol.com
Date: Sat, 10 Jun 1995 12:11:32 -0400
Subject: Re: Getting Users, etc.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is every bit as much a problem for the commercial laboratory as for the
academic scene. At Structure Probe, we have a multi-user, multi-operator
laboratory doing SEM and TEM for both biological and materials science
applications, including asbestos. In addition to all of the problems of
clutter and mess in any laboratory, we have the added risk of
cross-contamination. As the laboratory director, I can tell you that the
suggestions to get some "heavyweight" involved really don't help; giving
orders doesn't motivate the employees of the 1990's any better than it
inspires students or "crusty heads".

We are, however, having some success in appling the principles of Total
Quality Management (TQM) to this problem. TQM has been very successful for
us in reducing problems such as shipping the wrong product or the wrong
quantity of the right product to the customers of our SPI Supplies Division.
The same principles can work in any environment, including the laboratory.

Basically, we have let the people who use the laboratory work out their own
solution, with some guidance; this is one of the essentials of the TQM
approach. Their "solution" is to apply peer pressure to clean up, but on a
rotating basis, every laboratory employee is accountable to management for a
week for the cleanliness of the facilities. This is not the way I would have
directed that things be done, but because it came from the users (and
offenders) themselves, it works well.

I would be happy to dialog with anybody on the subject of quality management
in the laboratory environment.

Andy Blackwood

Andrew W. Blackwood, Ph.D.
Structure Probe, Inc.
63 Unquowa Road
Fairfield, CT 06430-5015
Phone: 1-203-254-0000
FAX: 1-203-254-2262
e-mail: AWBlackwoo-at-aol.com





From: allan.mitchell-at-stonebow.otago.ac.nz (Allan Mitchell)
Date: Mon, 12 Jun 1995 12:37:02 +1200
Subject: SEM verses TEM usage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ken and Terry,
Greatings from the deep south. We to were amalgamated about 5 years ago.
The technical staff of our Unit are funded from 3 Medical School
departments, 0.5 person from Micriobiology, 2.5 persons from Anatomy, 2
persons from Pathology.

Although housed in the Medical School we work for all the science
departments except the Dental School, ie Zoology, Botany, Physics,
Biochemistry, Marine Science, Geology etc.

3 have 3 TEM's and 1 SEM. It runs 6 to 10 per day, 5 to 6 days per week.
Most of the work is Zoology, Marine Science and Geology, very little
Medical School SEM users. There still seems to be a lot of fish and creepy
crawlies that need to be scanned but humans have been 'done to death'.

From our point of veiw ESEM or Cold stage SEM are the way to go. At the
conference here in September (I presume you know about our conference) will
be Brendon Griffin from Perth, Guru on environmental scanners and very
approachable (he has to be, we are paying for him to be here). Oxford
instruments are bringing an expert on Co;d stages I believe so come to the
conference to get the info.

If I can help with anything else just call

Regards

Allan



---------------------------------------------------------
Allan Mitchell
Senior Technical Officer
South Campus Electron Microscope Unit
C/- Department of Anatomy and Structural Biology
Otago Medical School

P.O. Box 913
Dunedin
New Zealand

Tel; National 03 479 7301 International 64 3 479 7301
Fax; National 03 479 7254 International 64 3 479 7254

"The Southernmost E.M. Unit in the World"






From: DVCCO-at-aol.com
Date: Mon, 12 Jun 1995 02:51:09 -0400
Subject: URL Web Site Video Camera Update

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


LM/EM URL Web Site Video Camera Guide Location Update!

***** After review of the web site feel free to address any questions to
((((((((((( DVC (DIRECTLY) via web site e-mail or dvcco-at-aol.com )))))))))))
***** PLEASE DO NOT TIE UP THE MICROSCOPY NET. Thank You

DVC Company wishes to introduce one of our many web sites to users of this
net, who wish to access 315876 bytes of detailed text, gifs, and performance
charts of the DVC video camera line.
------------------------------------------------------------------------
Our WWW URL Web Site is: http://www.edt.com/dvc/dvc.html
------------------------------------------------------------------------
Web files cover:
* DVC digital and analog cameras for cost effective sub-pixel accuracy
* PCI bus, Mac Nu-bus, Sun S-bus compatible board data
* Silicon Graphics (digital output) DVC cameras for Indy and Indigo 2!
* What to look for when choosing a CCD camera, (questions and answers) etc.
* Our goal is to present first hand information from the source. If you are
offended by information from whatever source, please hit your delete key now.

The complete HTML files detail:
DVC Monochrome CCD Digital/Analog Video Cameras operating at
real time video rates at 8 or 10 bits } 62dB signal to noise with no
cooling needed for extremely high sensitivity with many options.
Designed for quantitative microscopy and image analysis.
Models include:
DVC-10 RS-422 digital 10 bit with simultaneous RS-170 analog video
DVC-8 RS-422 digital 8 bit with simultaneous RS-170 analog video
DVC-0A RS-170 analog video/ ( upgradable to DVC-10/ DVC-8 models.)
This lets the researcher, cost effectively grow with the application.

* Compatible frame grabbers/ digitizers on the market with DVC offering
( the complete board, custom cable and our camera line, plug and play! )
for ( PCI bus, Mac Nu-bus, Sun S-bus, and the SGI digital out for the
Silicon Graphics Indy and Indigo 2 workstations! )

* Tunable Liquid Crystal Filters for:
* RGB sequential non real time 50ms switching from Red to Green to Blue
with no pixel alignment problems at a 50nm band width that can be
shifted by user for attachment to DVC or other monochrome cameras
or:
* Monochrome wavelength coverage from (400 - 1100nm), anywhere the user
wants to be within 50ms with 32 memories or sequential scan at any
user defined transition down to .1 nm step via RS-232 or TTL with any
user defined bandwidth from 5 to 50nm.

Request: When requesting info from DVC please offer us your:
* Complete address
* E-mail, phone, and fax
* Application, looking at what, on a microscope, low light req.
* Any special modifications, and time frame of project or need.
*** Down load as much information from our web site as possible.
DVC Company is proud to present information on a product manufactured in the
U.S.A. We send this web site update to you with the best of intentions and
trust it can be used as a guide to making the right decision when spending
precious funding on equipment. DVC realizes as our customers do, that your
camera is critical to your systems performance and not just incidental to it.





From: STEPHAN HELFER :      S.Helfer-at-rbge.org.uk
Date: Mon, 12 Jun 1995 09:58:43 BST
Subject: Re: Getting Users to Use Equipment Properly

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: Todd Voiles {tvoiles-at-unlinfo.unl.edu}

This may be an obvious thing to do for most of you, but as it has not been
mentioned before I risk stating the obvious: accountability is the key to a
multitude of multi-user problems. Keeping an accurate log book of users and
insisting in consistent log book entry, helps pinpointing problems (genuine
ignorance about procedures or blatant negligence). Must obviously be followed
up by personal contact with the miscreants. Initially it helps to give them the
benefit of the doubt; later on disciplinary measures may have to follow (e.g.
banning from the equipment or lab).

my tupence worth

Yours sincerely


Dr Stephan Helfer, SSO
Mycologist

Royal Botanic Garden Edinburgh, Inverleith Row, EDINBURGH EH3 5LR,
Scotland UK

email s.helfer-at-rbge.org.uk
phone: +44 (0)131 552 7171 ext 280
or +44 (0)131 459 0446-280 (direct digital VoiceMail line)
fax: +44 (0)131 552 0382




From: Biquan Lin :      blin-at-mcs.anl.gov
Date: Mon, 12 Jun 1995 08:13:41 -36803936 (CDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


UNSUBSCRIBE BLIN-at-MCS.ANL.GOV





From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Mon, 12 Jun 1995 09:20:12 -0400 (EDT)
Subject: Re: Getting Users to Use Equipment Properly

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We have a strict sign-up for microscopes, so we can identify who has used
a scope. The first time someone fails to take care of the equipment they
get a gently reminder from my staff, the second time they get a less
gentle reminder from me. Chronic problem causers are denied access to the
equipment. This final solution is not used often, but it has occurred.

That is how we have solved the problem.

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************

On Fri, 9 Jun 1995, Todd Voiles wrote:

} He at U of Nebraska we have been running into the re-occuring problem
} of users of the scopes and lab being overly messy and careless with
} the equipment. We've tried posting signs about it, writing standard and
} emergency operating proccedures to no avail. I'm sure other people
} have had similar problems in their labs........anybody want to share
} some solutions?
}
}
} Todd Voiles
} Facility Manager
} Central Facility for Electron Microscopy
} Center for Materials Research and Analysis
}
} University of Nebraska at Lincoln
}
} tvoiles-at-unlinfo.unl.edu
}
}




From: Michael Rock :      merock-at-u.washington.edu
Date: Mon, 12 Jun 1995 09:01:13 -0700 (PDT)
Subject: Re: Getting Users to Use Equipment Properly

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Todd-
try scheduling weekly clean-up sessions, include all users of the
facility, if they never seem to attend...restrict their access.
-Mike
On Fri, 9 Jun 1995, Todd Voiles wrote:

} He at U of Nebraska we have been running into the re-occuring problem
} of users of the scopes and lab being overly messy and careless with
} the equipment. We've tried posting signs about it, writing standard and
} emergency operating proccedures to no avail. I'm sure other people
} have had similar problems in their labs........anybody want to share
} some solutions?
}
}
} Todd Voiles
} Facility Manager
} Central Facility for Electron Microscopy
} Center for Materials Research and Analysis
}
} University of Nebraska at Lincoln
}
} tvoiles-at-unlinfo.unl.edu
}
}




From: Michael Boucher :      BOUCHER-at-tcrca.usbm.gov
Date: Mon, 12 Jun 1995 14:14:14 CST
Subject: Help:Frosted Petrographic Slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For all you petrographers; Help!

Our lab has for years been using frosted slides for polished thin sections.
Apparently only one manufacturer is left, Erie Manufacturing (Erie Scientific,
etc) and they are not satisfactory for our needs. All suppliers keep
sending these same slides. Ward's Scientific says that Erie is the
last remaining source for frosted petrographic slides. Making our own
will be very labor and time consuming. Any sources other than Erie?

The details are that the finish is very finely frosted, and there is
little "tooth" for the epoxy, so some of the phases easily pluck out
as we reach the end of the polishing stages. This began when we ran
out of our old supply from a now defunct company and began using the
new slides. Thanks for any help, advice, etc.
Please E-mail me direct unless you believe it would benefit others on
the server. Thanks
Mike
*****************************************************************
Michael L. Boucher Sr. Boucher-at-TCRCA.USBM.GOV
Geology-Mineralogy/Chemistry Labs Ph 612-725-4614
Twin Cities Research Center Fax 612-725-4527
U.S. Bureau of Mines Center 725-4500
Department of Interior
5629 Minnehaha Avenue South
Minneapolis, MN 55417-3099
U.S.A.
*****************************************************************





From: Fermin, Cesar :      fermin-at-tmc.tulane.edu
Date: 12 Jun 1995 14:33:26 -0600
Subject: Guidelines equip. usage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1409154719.35362-at-postoffice.tmc.tulane.edu}

Keeping a lab clean and avoiding abuses:

I identify well with this problem because we have here at Tulane Pathology
several core labs, one of which I direct. Remedy to this problem (disease) is
not readily available but there are pain relievers:

*1) The unit must have written guidelines of adherence with penalties for
abuse, training sessions for first time users and most importantly sharing of
the responsibility for keeping facility up by: a) users (do not work) or b)
someone else (see below),

*2) The manager must be willing, able, and have back up from superior to
enforce guidelines,

*3) Access to facility must be restricted by special combination locking
device or supervision to entrance of facility. In one of our facilities we
have a push-combination that is given after a signed Interdepartmental
Transation (IT) document is available,

*4) Users (or someone above them) must pay for service or generally users do
not give a ...,

*5) Supervisor must be able to stick to his gun, and realize that his position
is first keeping the facility and then make users happy. THERE IS NOT SECOND
WITHOUT THE FIRST.

*6) Do not ask for what it has not been explained. If succint guidelines are
not available, users are probably asked daily for something new they have not
heard.

*7) I am afraid that problems like those described on abuse of facilities
emmanates from our inability to tell others when and why they screw up: a)
first children (we praise evene when they done wrong), b) second teenagers
(they are encouraged after mediocraty), and c) third adults (demand to redo
what is wrong no matter how painful is for the asker and the doe)r. Example:
when one of my children breaks a glass as she learn to wash dishes I do not
praise that she is washing dishes. The first order to business is why the
glass broke and how not to do it again.

*8) When sticking to his guns, a supervisor will intially meet a lot of
unhappy faces, but if his main objective is to keep the facility runing and
clean and not to make friends, then I do not see what the problem is on
putting the foot down.

*9) Again, we are too afraid to tell people when they screw up and most of the
times we are unwilling to point out the terrible consequences of their lousy
work and complete lack of considerations for others. IT HAPPENS THAT TO DO
THE RIGHT THINGS DO NOT ENTER TO BRAIN BY OSMOSIS!

*10) Bottom line: My Dads uneducated moto: If they do not give a ...,
neither do I, and of course my boss must agree to it too!

I hope that this help,

******************************************************************
*Cesar D. Fermin, Ph.D \|T|/ Fax (504) 587-7389 *
*Tulane Medical School /|M|\ Voice Mail (504) 584-2618 *
*Pathology/SL79 \|C|/ Secretary (504) 584-2436 *
*New Orleans La 70112-2699 /|*|\ Lab/Techn. (504) 584-2521 *
* {Fermin-at-tmc.tulane.edu} Dir. Morphological Services *
******************************************************************





From: BARRY PYLE :      umbbp-at-msu.oscs.montana.edu
Date: Mon, 12 Jun 1995 13:56:31 MDT
Subject: Epifluorescence

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


June 12, 1995

Does anyone in the U.S. know which company(ies) market Citifluor mounting fluid
for epifluorescence microscopy?




From: Patty Jansma :      plj-at-manduca.neurobio.arizona.edu
Date: Mon, 12 Jun 1995 13:37:18 -0700 (MST)
Subject: Re: Epifluorescence

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ted Pella carries Citifluor mounting media.(cat. 19470).
P.O. Box 492477
Redding, CA 96049

Patty Jansma Tel:602-621-6671
plj-at-manduca.neurobio.arizona.edu
Arizona Research Labs Division of Neurobiology
University of Arizona


On Mon, 12 Jun 1995, BARRY PYLE wrote:

} June 12, 1995
}
} Does anyone in the U.S. know which company(ies) market Citifluor mounting fluid
} for epifluorescence microscopy?







From: dlmedli-at-california.sandia.gov (medlin douglas l)
Date: Mon, 12 Jun 1995 15:50:25 -0700
Subject: file transfer:TN 5500 to Macintosh

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have an old Tracor Northern 5500 EDS analysis system from which
we are trying to transfer files (through the serial port) to a Macintosh
using the Xi program on the 5500 and Versaterm on the Mac. Has anyone had
similar experience (e.g. I'm interested in the proper pin connections and
software configurations, information our Noran service reps have not
been able to provide). Any help would be appreciated.

+---------------------------------------------+
! Douglas L. Medlin !
! Physical Properties of Materials Department !
! Mail Stop 9402 !
! Sandia National Laboratories !
! Livermore, California 94551 !
! !
! (510) 294-2825 !
! dlmedli-at-california.sandia.gov !
+---------------------------------------------+




From: Jane A. Fagerland (708) 935-0104 :      FAGERLAND.JANE-at-igate.pprd.abbott.com
Date: Mon, 12 Jun 1995 19:01:00 -0600 (CST)
Subject: slobs in the lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta RANDB; Relayed; Mon, 12 Jun 1995 19:55:21 -0600
Mr-Received: by mta MCM$RAND; Relayed; Mon, 12 Jun 1995 19:55:22 -0600
Mr-Received: by mta RANDD; Relayed; Mon, 12 Jun 1995 19:55:29 -0600
Alternate-Recipient: prohibited
Disclose-Recipients: prohibited
Content-Return: prohibited

A few thoughts.....

Just as there are lots of different types of people in this world, there
are lots of different reasons for bad lab manners. And no one right way
to solve the problem.

In my experience, a heavy-handed approach works with very few people.
More often, it brings out some very creative passive/aggressive
responses and a lot of resentment. While I agree that some "unhappy
faces" are inevitable, much unpleasantness can be avoided by treating
people respectfully. Some people are simply unaware that they are making
messes or creating problems for those who come after them - these people
are easily "retrained" by telling them what the standards are.

The infuriating people are those who really don't care and who will not
respond to civil requests for them to change their lab habits. They
make life miserable for all around them, and ultimately, must be
banished from the lab.

However, I take exception to the idea that keeping a facility running
efficiently and keeping users on friendly terms are mutually exclusive.
I think a "kinder, gentler" approach is the best default mode and that
the Eastwood/Schwarzenegger approach should be reserved for those people
who don't respond to forthright requests. (This is not to say that I
haven't personally wanted to remove body parts from some recalcitrant
slobs, just that this is not always the best way to solve the problem!)

Jane A. Fagerland
Dept. Cellular and Microscopic Research
Abbott Laboratories
Abbott Park IL 60064






From: Dr H. Cama :      hc203-at-cus.cam.ac.uk
Date: Tue, 13 Jun 1995 09:08:33 +0100 (BST)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please include my e-mail address in your mailing list.

Thanks.

H. Cama
University of Cambridge
UK




From: rsartore-at-arl.mil (Sartore, Richard G.)
Date: Tue, 13 Jun 1995 08:12 -0500 (EST)
Subject: Re: file transfer:TN 5500 to Macintosh

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have successfully tranferred image and xray map file using Xi from the
TN5500 to
a PC and SUN station over the serial port. It is slow. A 1M image file
takes about an hour.
Transfer is in text or ASCII mode. The Tracor file format must be
translated to a more standard
format for display. A West Point Cadet, during summer visit, wrote a
Pascal/C routine to convert Tracor Northern file to PGM format for
display on monitor. The image was also
successfully printed on laser printer by converting PGM file to a
PostScript file.
Connector was wired for standard RS-232 tranfer mode.
Can get in touch with Cadet for availability of code and connector
wiring. Please let me know.
Richard Sartore
US Army Research Laboratory
AMSRL-PS-DC
Fort Monmouth, NJ 07703
908-427-2261
rsartore-at-ftmon.arl.mil






From: STEELE-at-KRDC.INT.ALCAN.CA
Date: Tue, 13 Jun 1995 08:22:51 -0500 (EST)
Subject: TEM HELP: PTA procedure request

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Would someone in the know have a minute to post a simple protocol
for negative staining with PTA?

I have a suspension of polymer "particulate", and wish to get a measure of
the size distribution. I plan to disperse onto a formvar substrate, then
shadow with Pt/C to get a basic idea of size, but would also like to try
negative staining to obtain images suitable for image analysis.

Comments would be welcome.

Thanks in advance.





From: JOHNA-at-SCI.WFEB.EDU
Date: Tue, 13 Jun 1995 09:45:20 -0400 (EDT)
Subject: Re: TEM HELP: PTA procedure request

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I usually use a 2% or 4% solution of PTA (aqueous) and adjust the pH to 6.9
or 7.4 (specimen dependent) with KOH. The actually negative staining is as
is typical when using UA.

Good luck

John

___________________________________________________
| |
| John G. Aghajanian, Ph.D. |
| Worcester Foundation for Experimental Biology |
| 222 Maple Avenue |
| Shrewsbury, MA 01545-2737 |
| |
| Tel: 508 842-8921 ext. 147, 161 |
| Fax: 598 842-9632 |
| JOHNA-at-sci.wfeb.edu |
| |
|_________________________________________________|





From: Facility Manager (Todd Voiles) and
Date: Tue, 13 Jun 1995 10:26:09 -0400
Subject: All the responses to "Getting users to Use equip"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506131525.AA14838-at-unlinfo.unl.edu}
X-Sender: tvoiles-at-129.93.1.11
X-Mailer: Windows Eudora Version 1.4.4
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

WOW,

This must be a real sore point for alot of managers out there, it's good
to know I'm not alone in my misery.

After 31 responses (a vertitable deluge for my underused account) the
general consensus is that I need to grow some teeth and wear some
bigger boots....steel toed would be appropriate.

One responder put it perfectly......"Most good managers learn to be
Tough n' Nice not Nice n' Tough.

Thanks for all the good advice.....I've finished writing a new use policy for
our lab (which follows at the bottom). I would like some feedback on it.....
is it to nice, to mean, uninforcable?

Anyone can feel free to use any or all of it for your own purposes (for a
small copyright fee of course ; ) }.

Thanks.

What follows is the proposed new use policy for our laboratory-




Memorandum

To: Persons using the Central Facility for Electron Microscopy
and its equipment


Re: The new policy concerning the Specimen Preparation room

Due to the current state of Specimen Preparation room (a mess)
a new policy has been implemented and will go into effect at the
end of June when the policy is finished and avialable for review.
All persons wishing to use the specimen preparation room and its
equipment will first have to notify Facility Manager so he can
detail this new policy. After the user agrees to the new policy he
or she is free to use the equipment in the appropriate manner.

**Any persons found in the specimen preparation room without first
having talked to Facility Manager will be expelled and lab
privileges revoked until further notice.**


The new policy is as follows:


Samples:

*All samples or materials brought into the laboratory or used by
users must be kept in their respective drawers or areas at all
times when not being actively being worked on and an MSDS is required
to be on file at all times.

*Any samples left unattended and unmarked for more than 2 hours may
be seized by Facility Manager and held until picked up by the user they
belong to. This action will be deemed a violation.


Equipment Use:

*All users wishing to use a specific piece of equipment will need
to be trained in its operation by the Facility Manager or one of
several officially appointed student instructors.

*All users will follow either the manufacturers manual or the
instruction sheets provided by Facility personnel. Deviation
from these instructions will be deemed a violation.

*All users must fill out use logs for equipment when operating it,
failure to do so will be deemed a violation.

*Any users damaging or finding equipment damaged or any lack of
supplies should contact the Facility Manager immediately. Failure
to report damage will be deemed a violation.

Consequences of Violation:

*Upon the first violation a warning will be issued and the user
will be restricted to working only during normal lab hours (8-5).

*Upon the second instance an additional warning will be issued and
the user will be restricted to working only when Facility Manager
is present in the laboratory and a notice will be sent to their
advisor and the Facility Director.

*Upon the third violation all laboratory privileges will be revoked
and a meeting with the user and their advisor and the Facility
Director will need to be arranged to resolve the problem.


These restrictions will be lifted when the user sufficiently
demonstrates that he or she understands the problem and works to
solve it.


A full copy of the new policy is in development and will be made
available in the Facility when it is completed (around the end of
June). If you have any questions regarding this policy or its
elements please contact the Facility Manager (Todd Voiles) at
2-8762 or tvoiles-at-unlinfo.unl.edu.




Todd Voiles
Facility Manager
Central Facility for Electron Microscopy
Center for Materials Research and Analysis

University of Nebraska at Lincoln

tvoiles-at-unlinfo.unl.edu





From: Greg :      GREG-at-umic.umic.sunysb.edu
Date: Tue, 13 Jun 1995 11:52:48 EST5DST
Subject: Lab Mess

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This sounds like a good way to go, Mr. Voiles.
If the policy is followed, word will spread and the abuse
should go away. I work at a core facility and "no one" is
left unsupervised and there is no mess.

Good luck!

Gregory Rudomen
University Microscopy Imaging Center
S.U.N.Y. at Stony Brook
Greg-at-umic.umic.sunysb.edu





From: BARRY PYLE :      umbbp-at-msu.oscs.montana.edu
Date: Tue, 13 Jun 1995 10:25:19 MDT
Subject: Epifluorescence - Thanks!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


June 13, 1995

Thanks to all who sent info on sources of Citifluor and Biomount for
epifluorescence microscopy. It's great to get information like this so easily!

Barry Pyle, Department of Microbiology, Montana State University-Bozeman




From: Beverly E Maleeff :      Beverly_E_Maleeff%notes-at-sb.com
Date: 13 Jun 95 12:55:28 EDT
Subject: Multi-User EM Facilities...again

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Having seen the messages about EM facility etiquette, or the lack thereof, I'd
like to remind everyone of a session at the upcoming MSA meeting that may be of
interest.

The MSA Technologists' Forum (TF) sponsors a roundtable discussion each year at
the Annual Meeting. This year's topic is, "EM Facility Management: Survival
Part III - What Works and What Doesn't". It is a continuation of the
discussions held at the last two meetings, and the TF offers a forum (pun
intended) for just this kind of ongoing dialogue, be it at this session, or at
our exhibit booth throughout the meeting week. If you plan to be in Kansas
City, I invite you on behalf of the TF to join us and bring these kinds of
concerns for discussion. More details about the session will be available in
the MSA meeting program.

Regards,
Bev Maleeff, TF Chairman-Elect

SmithKline Beecham Pharmaceuticals
Toxicology-US, UE0462
709 Swedeland Road
King of Prussia, PA 19406
610/270-7987
610/270-7202 fax
maleeffbe-at-sb.com or Beverly_E_Maleeff%Notes-at-sb.com





From: Michael Rock :      merock-at-u.washington.edu
Date: Tue, 13 Jun 1995 11:16:53 -0700 (PDT)
Subject: Re: Controlling People in User Laboratories

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree!
as a microscopy lab manager we (myself included) have the wonderful
opportunity to open the eyes and imaginations of bright eager young
people. most likely the equipment is not our own, and is usually covered by
service contract, lighten up. students will make mistakes (just like us,
who have years...decades of experience) . this philosophy in no way
gives them the right to rage wrecklessly through the lab, but few that I
work with ever do that. most of the "accidents" were just that, and I
take most of the responsibility for not preparing the students for the
possible outcomes. keep your sense of humor life is short. teach the
student how to avoid damaging the equipment (operator manuals &
individual instruction), encourage them to experiment, and employ
the appropriate security for the facility.
be a teacher, not a policeman!
-Mike Rock
U.W. Zoology Dept.

On Tue, 13 Jun 1995 JMARDINLY-at-IMO.intel.com wrote:

} My sympathies to those of you out there who have no sense of humor. It
} is important to have a sense of humor when managing user laboratories. For
} those who replied to my suggestion of hiring Arnold Schwartzenegger and
} Clint Eastwood as special assistants as too heavy handed, and did not recognizeit as sarcasm, get a life. Managing a user laboratory is akin to managing
} the human condition, and that is a most challenging task. You will have a
} hard time stopping people from being themselves, and you cannot stop people
} from inventing insane new ways to misuse equipment. Just teach the best you
} can and grin and bear it when the results come out different from what you had
} hoped for.
} John Mardinly
} Intel Materials Technology
}




From: STEELE-at-KRDC.INT.ALCAN.CA
Date: Tue, 13 Jun 1995 08:22:51 -0500 (EST)
Subject: TEM HELP: PTA procedure request

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {1995Jun13.111734.1158885706-at-ms.sjdccd.cc.ca.us}
To: MICROSCOPY-at-aaem.amc.anl.gov (microscopy list),
STEELE-at-krdc.int.alcan.ca (STEELE)

Depending on exactly what type of polymer it is you might want to stain it
with ruthenium tetroxide or osmium tetroxide if the neg stain does not
work.

When using the PTA: You might also want to try to adjust the pH of the PTA
to 5.0 as sometimes PTA works as a negative or positive stain depending on
exactly what pH and ionic condition the specimen is
in.
_______________________________________________________________________________



Would someone in the know have a minute to post a simple protocol
for negative staining with PTA?

I have a suspension of polymer "particulate", and wish to get a measure of
the size distribution. I plan to disperse onto a formvar substrate, then
shadow with Pt/C to get a basic idea of size, but would also like to try
negative staining to obtain images suitable for image analysis.

Comments would be welcome.

Thanks in advance.



______________________________________________________________________________
San Joaquin Delta College World Wide Web:http://www.sjdccd.cc.ca.us
5151 Pacific Avenue email:webmaster-at-ms.sjdccd.cc.ca.us
Stockton, CA 95207
general information:(209) 474-5151 or FAX-at-(209)474-5600





From: Michael Rock :      merock-at-u.washington.edu
Date: Tue, 13 Jun 1995 15:07:05 -0700 (PDT)
Subject: magneto optical disc reader

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


i am looking for a lab in the great northwest who has a SONY magneto
optical disc reader. one of our researchers went to san diego, aquired
some confocal data, and copied it on a SONY worm disk.
we unfortunately for her, have a panasonic worm dirve. the disk holds
approx. the same amt of data, but will not fit into the drive. i need to
down load this data, does any one here inseattle have a SONY worm drive ?
if so please contact me soon.
tia
merock-at-u.washinton.edu




From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Tue, 13 Jun 1995 17:44:52 -0600
Subject: EPMA: Fly ash trace elements

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v01510102ac03d2876366-at-[131.230.97.56]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear Analysts-
One of our researchers wishes to map trace elements found in fly ash.
Anyone have any good references or protocols they would share with us? I
assume that the samples should be enrobed in resin, polished, etc. but
details would be appreciated. Thank you very much.

John J. Bozzola
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
Phone: 618-453-3730
Fax: -2665
Email: bozzola-at-siu.edu OR bozzola-at-qm.c-cem.siu.edu






From: bjg-at-uniwa.uwa.edu.au (Brendon J. Griffin)
Date: Wed, 14 Jun 1995 10:05:59 +0800
Subject: What is the ideal EM centre profile

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,

At the University of Western Australia we have a Centre for Microscopy and
MIcroanalysis. Currently we have 3 TEMs, 5 SEMs (inc FESEM, ESEM) and a
confocal together with a range of spec prep facilities. We run quarterly
2-3 day short courses in all aspects of EM and confocal. We have a Director
and 2 academics (covering materials, geoscience and biological), a
secretary and 5.5 technical support staff. Users drive the instruments
themselves following successful attendance at one of the training courses.
There are currently no charges as we are funded by a group of Faculties -
the budget is minimal and ~30% is raised by consulting and competitive
grants by the academics. Major equipment for the last five years has been
gained through national competitive funding with seed funding from the
University. We also act in a regional capacity supporting EM research from
two other local Universities.

As usual we are under audit.

I would appreciate greatly any comment on:

ideal and actual centre profiles - what equipment is actually needed for a
cross-campus or single discipline facility

funding strategies - charge rates : do you charge? what does it cost to
charge? is the charging a full cost recovery? where does the rest come
from?

staff profile - actual and ideal: particularly academic or technical or
mixture - note here this may be historically controlled, ie if you have
academics in a centre then the user groups may not develop the expertise
themselves

equipment replacement policies - is there a plan or is it serendipidous as
opportunities arise?

equipment maintenance policies - eg in-house or service contract



I am very concerned at the difficulties many EM facilities are having
world-wide and plan to try and present a review of trends apart from using
the data ourselves.

All data will be kept in confidence w.r.t. location/people etc but I would
need some ID to verify original data.


If I get enough feedback I will put a draft/summary back through Nestor.
Thanks in advance

Happy kangaroos


Brendon
Brendon J. Griffin
Centre for Microscopy and Microanalysis
The University of Western Australia
Nedlands, WA, AUSTRALIA 6907
ph 61-9-380-2739 fax 61-9-380-1087





From: dietmar.reiter-at-uibk.ac.at (Dietmar Reiter)
Date: Wed, 14 Jun 1995 12:09:38 +0100
Subject: Imaging gray cataract dimmed eye lenses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear bio-medical Imagers,

any experiences with confocal imaging of cataract lenses to get a
representation of the blurred area in 3D? Is confocal imaging possible at
all?

Thanks for hints, -Dietmar-


+++ Dietmar Reiter, Dept. of Zoology and Limnology
+++ University of Innsbruck
+++ Technikerstrasse 25, A-6020 Innsbruck, Austria
+++ phone: (43)-512-507 ext. -6170, fax ext. -2930






From: EMLAB-at-opus.mco.edu
Date: Wed, 14 Jun 1995 08:52:47 -0500 (EST)
Subject: Re: TEM HELP: PTA procedure request

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






From: mrc-at-vax.ox.ac.uk
Date: Wed, 14 Jun 1995 14:30:36 +0100
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: mrc-at-vax.ox.ac.uk





From: sulovsky-at-sci.muni.cz (Petr Sulovsky)
Date: Wed, 14 Jun 1995 10:18:09 -0400
Subject: CamScan phone/e-mail?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506140820.AA21400-at-elanor.sci.muni.cz}
X-Sender: sulovsky-at-elanor.sci.muni.cz
X-Mailer: Windows Eudora Version 1.4.3
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hello Microscopists,

could someone help me with current phone number of CamScan (Mr. Holliday's,
preferably), or (better), their e-mail code (they once had some)?

Thanks

Dr. Petr Sulovsky
Dept. of Mineralogy, Petrology and Geochemistry
Faculty of Science
Masaryk University, Brno
Postal address: Kotlarska 2
CZ 611 37 Brno
Czech Republic
Fax +42-541211214, phone +42-541129231
E-mail: sulovsky-at-sci.muni.cz





From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Wed, 14 Jun 1995 14:38:05 +1100
Subject: Image processing tutorial software?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: st004718-at-brandywine.otago.ac.nz
Message-Id: {v01510101ac0407111404-at-[139.80.120.172]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


Does anyone know of any software package which is designed to introduce
people to digital image processing, perhaps a sort of tutorial which would
guide one through some of the basic tools such as filters, edge
enhancement, equalisation, contrast manipulation, etc?
Thanks in anticipation,
Richard E

Richard Easingwood
South Campus Electron Microscope Unit
Otago Medical School
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

"The southernmost electron microscope unit in the world"






From: cbarba-at-eucmax.sim.ucm.es (andy johnson)
Date: Wed, 14 Jun 1995 18:15:37 +0100
Subject: Image processing tutorial software?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe






From: Wan Junzuo :      junzuo-at-mcmail.CIS.McMaster.CA
Date: Wed, 14 Jun 1995 12:43:42 -0400 (EDT)
Subject: unsubsribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I would like to unsubscribe, please.

junzuo wan




From: diamond-at-chem.psu.edu (John V. Badding)
Date: Wed, 14 Jun 1995 12:41:56 -0400
Subject: unsubsribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506141640.MAA18871-at-portal.chem.psu.edu}
X-Sender: diamond-at-mailhost.chem.psu.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Can anyone advise me where we could go to perform EELS measurements on a
TEM on some carbon samples? preferably within several hours drive of Penn
State in central Pennsylvania. We have done these measurements on a STEM,
but want to have better simultaneous diffraction and EELS capabilities.
Thank you.
John Badding






From: GVKM07A-at-prodigy.com (DR CHARLES A GARBER)
Date: Tue, 13 Jun 1995 23:40:28 EDT
Subject: EPMA: Fly ash trace elements

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

On June 13, John Bozzola wrote about fly ash particles:

1] We have gotten or best EDS data on thin sections via TEM. We use
our SPI Supplies "Materials Science" diamond knife and our SPI-Pon 812
Resin system. The sections seem also to be suitable for RBS analysis.
Vacuum embedding is important and then you have to be careful that the
largest particles are not "pulled out" of the section producing a non-
representative picture of the sample as a whole.

2] If a TEM is not going to be used, then mounting the particles using
our "Tacky Dot" slides and then embedding and polishing down to cross
sections would give you a nice orthogonal array of cross sectioned and
polished particles which makes the EPMA a "breeze" compared to what
would otherwise be the case.

Chuck

Charles A. Garber, Ph. D.
PRESIDENT
STRUCTURE PROBE, INC.
PO Box 656
West Chester, PA 19381-0656

Ph: (610) 436-5400
FAX:(610) 436-5755
e-mail: GVKM07A-at-prodigy.com





From: JOHNA-at-SCI.WFEB.EDU
Date: Wed, 14 Jun 1995 11:46:08 -0400 (EDT)
Subject: LR Gold

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506141834.AA28863-at-riker.ml.wpafb.af.mil}

G'Day Folks,

I have been using LR White for years, both for straight morphology and
immunocytochemistry, but have not used LR Gold. I know someone out there
is using LR Gold for light & EM immunocytochemistry since I seem to
remember reading something about it in this forum.

I'd appreciate it if you folks could answer a few questions. I'd like to
give LR Gold a try for light and EM IMC and in situ hybridization.

1. If using lightly fixed tissue (4% PA + 0.1-0.2% GA) does need
to add PVP to the methanol? Can you use ethanol instead of methanol? The
infiltration times listed in the LR Gold brochure are quite long and
haven't changed since they first started selling the stuff; is it that
difficult to get this stuff into tissues and cultured cells or were they
just being extremely cautious when they wrote up the protocol.

2. Can L.R. Gold be heat polymerized in an oven?

3. Any other helpful hints?

Thanks very much in advance. See you in K.C.

John

___________________________________________________
| |
| John G. Aghajanian, Ph.D. |
| Worcester Foundation for Experimental Biology |
| 222 Maple Avenue |
| Shrewsbury, MA 01545-2737 |
| |
| Tel: 508 842-8921 ext. 147, 161 |
| Fax: 598 842-9632 |
| JOHNA-at-sci.wfeb.edu |
| |
|_________________________________________________|





From: Michael Boucher :      BOUCHER-at-tcrca.usbm.gov
Date: Wed, 14 Jun 1995 14:29:10 CST
Subject: NORAN Voyager (WDS?) users group

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Another user recently contacted me by e-mail about problems with his
Voyager system, describing exactly my recent problems with it too.
I would like to get a e-mail/phonel list of Voyager users, but particularly
WDS users, since that system is still in development and version 3 is
due out (to be delivered to me anyway) in late August. Please e-mail
me directly if you are interested in starting up a "user's group"
that would share help and advice by phone and e-mail; not burdening
the microscopy listserver with our traffic unless it seems to be of
general interest or new people ask for help regarding Voyager. Any
takers?
Mike
*****************************************************************
Michael L. Boucher Sr. Boucher-at-TCRCA.USBM.GOV
Geology-Mineralogy/Chemistry Labs Ph 612-725-4614
Twin Cities Research Center Fax 612-725-4527
U.S. Bureau of Mines Center 725-4500
Department of Interior
5629 Minnehaha Avenue South
Minneapolis, MN 55417-3099
U.S.A.
*****************************************************************





From: bjg-at-uniwa.uwa.edu.au (Brendon J. Griffin)
Date: Wed, 14 Jun 1995 15:21:55 +0800
Subject: What is the ideal EM centre profile

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506140713.PAA26276-at-uniwa.uwa.edu.au}

Dear all,

At the University of Western Australia we have a Centre for Microscopy and
MIcroanalysis. Currently we have 3 TEMs, 5 SEMs (inc FESEM, ESEM) and a
confocal together with a range of spec prep facilities. We run quarterly
2-3 day short courses in all aspects of EM and confocal. We have a Director
and 2 academics (covering materials, geoscience and biological), a
secretary and 5.5 technical support staff. Users drive the instruments
themselves following successful attendance at one of the training courses.
There are currently no charges as we are funded by a group of Faculties -
the budget is minimal and ~30% is raised by consulting and competitive
grants by the academics. Major equipment for the last five years has been
gained through national competitive funding with seed funding from the
University. We also act in a regional capacity supporting EM research from
two other local Universities.

As usual we are under audit.

I would appreciate greatly any comment on:

ideal and actual centre profiles - what equipment is actually needed for a
cross-campus or single discipline facility

funding strategies - charge rates : do you charge? what does it cost to
charge? is the charging a full cost recovery? where does the rest come
from?

staff profile - actual and ideal: particularly academic or technical or
mixture - note here this may be historically controlled, ie if you have
academics in a centre then the user groups may not develop the expertise
themselves

equipment replacement policies - is there a plan or is it serendipidous as
opportunities arise?

equipment maintenance policies - eg in-house or service contract



I am very concerned at the difficulties many EM facilities are having
world-wide and plan to try and present a review of trends apart from using
the data ourselves.

All data will be kept in confidence w.r.t. location/people etc but I would
need some ID to verify original data.


If I get enough feedback I will put a draft/summary back through Nestor.
Thanks in advance

Happy kangaroos


Brendon
Brendon J. Griffin
Centre for Microscopy and Microanalysis
The University of Western Australia
Nedlands, WA, AUSTRALIA 6907
ph 61-9-380-2739 fax 61-9-380-1087





From: rybicka-at-acsu.buffalo.edu
Date: Wed, 14 Jun 1995 12:32:18 -0400
Subject: glycogen ultrastructure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All Those Interested in Glycogen,

There is a short review in MICROSCOPY TODAY, October, 1994 entitled
"GLYCOGEN GFRANULES REVISITED". It explains that glycogen in the tissue
contains covalently bound protein. The event is well established in
biochemistry, but still misinterpreted in EM. Since this protein contains
enzymes involved in glycogen metabolism, the structures composed of
glycogen-protein complex were considered as cellular organelles and called
glycosomes in 1968 ! So called "glycogen granules" stainable with uranium
and lead represent protein component (enzymes) attached to glycogen.
Glycogen itself remains invisible after routine staining with uranium and
lead salts because it does not form ionic bonds. Apparent removal of
"glycogen granules" by uranyl acetate (UA) or other acids used before
tissue dehydration, is in fact the removal of protein component. It appears
because the bond of protein to glycogen is sensitive to the change in pH
(traditional purification of glycogen is performed by the treatment with
strong alkali or acids). The removed soluble protein is presumably washed
out during dehydration. Glycogen appears as small (3 nm) particles which
normally form (20-30nm) aggregates attached to protein. Glycogen is not
fixed per se, but it is stabilized due to the fixation of the bound
protein. After removal of protein by acidic treatment, glycogen particles
floate freely in the cell forming large irregular clumps. In routinely
stained tissue these clumps appear as white spots which may even suggest a
poor plastic polymerization. However, histochemical staining (Thiery
technique) shows that these spots are composed of the clumps of 3 nm
particles of glycogen. First EM study of the subject was: RYBICKA, K.
1979. Virchows Archiv B. Cell Pathol. 30, 355-47. Other references are
included in the mentioned review in MICROSCOPY TODAY. If you cannot get it,
please, let me know, and I will write you more details and references.

Considering present status of biochemical knowledge about glycosomes,
microscopic and molecular biology techniques seem to offer a good field for
further research.

Good luck,

Krystyna Kielan Rybicka
SUNY at Buffalo
Physiology / Neurobiology
Buffalo, NY

---
-------------------------------------------------------------------------
email address rybicka-at-acsu.buffalo.edu
phone number (lab) 716 829 3575
-------------------------------------------------------------------------





From: EvexAnalyt-at-aol.com
Date: Wed, 14 Jun 1995 19:35:19 -0400
Subject: Re: Image processing tutorial software?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Yes We have available for you a demo version of ImagePro please respond to

Evex Analytical Instruments
PO Box 630
Princeton, NJ 08542
USA
(908)874-3800
EvexAnalyt-at-aol.com




From: EvexAnalyt-at-aol.com
Date: Wed, 14 Jun 1995 19:43:16 -0400
Subject: Re: EDS: file transfer from a TN5500 to a Mac

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Evex Analytical Instruments has developed the VIDX INTERFAZER

The VIDX INTERFAZER comes in 2 models the operate on a PC.

One unit is transfers the image instanly-live to the PC for low resolution
Images or Dot Mappings

The other uses transfers the images & Spectra at a slower pace with greater
resolution.

For more information please contact

Evex Analytical Instruments
PO Box 630
Princeton, NJ 08542
(908)874-3800




From: EvexAnalyt-at-aol.com
Date: Wed, 14 Jun 1995 19:44:30 -0400
Subject: Re: EDS: file transfer from a TN5500 to a PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Evex Analytical Instruments has developed the VIDX INTERFAZER

The VIDX INTERFAZER comes in 2 models the operate on a PC.

One unit is transfers the image instanly-live to the PC for low resolution
Images or Dot Mappings

The other uses transfers the images & Spectra at a slower pace with greater
resolution.

For more information please contact

Evex Analytical Instruments
PO Box 630
Princeton, NJ 08542
(908)874-3800




From: EvexAnalyt-at-aol.com
Date: Wed, 14 Jun 1995 19:51:46 -0400
Subject: Tracor/Noran Kevex, PGT USER'S GROUP Q & A

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please e-mail me directly if you are interested in a USER'S GROUP
that would share advice by phone and e-mail





From: MicroToday-at-aol.com
Date: Wed, 14 Jun 1995 22:54:44 -0400
Subject: Manufacturers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Group -
As an "ex-manufacturer" I am particularly interested in the "aroma" presented
about manufacturers as carried in this media. As an example I refer to a
recent note regarding the Noran/TN 5500 by Douglas Medlin at Sandia which
read "I'm interested in the proper pin connections and software
configurations, information our Noran service reps have not been able to
provide".
Since I happened to had a "responsible" position at TN when this system was
sold, I looked into the matter. The facts are that the pin connections were
supplied to Sandia some 3 months ago. And, with Noran's offer to fax another
set, they found the previous set. Also - a very qualified Noran service tech
(from the home office) recently spent 3 days on-site checking out the system
- with the conclusion that the problem lies in the Mac PC.
Two points in conclusion:
1) I suggest that participants in this listserver be very careful in their
reference to manufacturers or suppliers - be they positive or negative. The
reason being that such, from folks like you all, do logically tend to carry
much weight.
2) I happen to be a bit proud of the work/products/service that TN provided
when I was there. And I have no comment on the current support, etc. of any
of the EDS companies.
Cheers,
Don Grimes, Microscopy Today




From: diamond-at-chem.psu.edu (John V. Badding)
Date: Thu, 15 Jun 1995 08:53:40 -0400
Subject: EELS on the East Coast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I sent this message out previously without any subject. I apologize for
the repeat for those of you who have already read it.

Can anyone advise me where we could go to perform EELS measurements on a
TEM on some carbon samples? preferably within several hours drive of Penn
State in central Pennsylvania. We have done these measurements on a STEM,
but want to have better simultaneous diffraction and EELS capabilities.
Thank you.
John Badding






From: CHAFFEYN :      NIGEL.CHAFFEY-at-bbsrc.ac.uk
Date: Thu, 15 Jun 1995 09:36:32 -0500
Subject: Digitising pad

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X400-Received: by mta bottom.ctd.anl.gov in /PRMD=ANL/ADMD= /C=US/; Relayed;
Thu, 15 Jun 1995 08:37:19 -0500
X400-Received: by /PRMD=ESnet/ADMD= /C=US/; Relayed;
Thu, 15 Jun 1995 08:37:17 -0500
X400-Received: by /PRMD=uk.ac/ADMD= /C=gb/; Relayed;
Thu, 15 Jun 1995 08:37:12 -0500
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Thu, 15 Jun 1995 09:36:21 -0500
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Thu, 15 Jun 1995 09:36:23 -0500
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Thu, 15 Jun 1995 08:42:09 -0500
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; converted (ia5 text (2)); Relayed;
Thu, 15 Jun 1995 09:36:32 -0500

Dear All,

Can anybody help a colleague of mine? He is trying to locate a source
for a 'Calcomp 2000 digitising pad and cursor' to fit a VIDS II image analysing
system on an Apple IIe. The previous pad plus cursor has disappeared, but was
obtained from Analytical Measuring Systems Ltd of Saffron Walden, Essex (UK).
He has been unable to trace the original company. Therefore:
can anybody suggest a source for the required item?
does anybody have one that is surplus to their requirements, which
could be 'acquired', or,
can anybody tell me if the original company is trading under another
name?

Many thanks in advance

Nigel.Chaffey-at-BBSRC.AC.UK..




From: emccjb-at-ttuhsc.edu (Charles J. Butterick)
Date: Thu, 15 Jun 1995 09:36:32 -0500
Subject: MSA meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: MICROSCOPY-at-AAEM.AMC.ANL.GOV

Greetings,
Would someone please send MSA's address, phone number, email and/or
fax so that I can get details on August's meeting? Thanks.



Charles J. Butterick (Chuck)
Electron Microscopy Center
Department of Cell Biology
and Biochemistry
Texas Tech University Health
Sciences Center
3601 4th Street
Lubbock, Texas 79430

vox (806) 743-1633
fax (806) 743-1219
email emccjb-at-lubb.ttuhsc.edu or
chuck-at-micron1.lubb.ttuhsc.edu






From: Eugene Bluhm :      bluhm-at-umr.edu
Date: Thu, 15 Jun 1995 10:39:39 -0600 (CDT)
Subject: MSA meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Thu, 15 Jun 1995 11:38:41 -0400 (EDT)
Subject: EM AND LM IMMUNOSTAINING: Ab to FITC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have a client who wishes to immunostain for FITC. He would like to do
this at the light microscope level in both parafin and fixed,
cryosectioned material. He would also like to do this at the EM level by
pre-embed immunostaining.

My questions are:

1. Any recommendation on which of the Anti-FITC anti-bodies work well (or
don't work) for immunhistochemistry? [species the antibody was raised in
is not limiting at this point]
a. Parafin embedded material?
b. Crosections?
c. Fixed, but unembed material?

2. Can anyone provide a good reference (I have a couple of papers but the
methods are sketchy) or protocol for any of the three above staining
situations with Anti-FITC? Concentrations, times, etc?

3. Are there any unusual pitfalls to worry about? We have considerable
experience with immunostaining, so have the basics down.

I appreciate any help y'all can provide. Thanks

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************





From: sbianchi-at-dbag.unifi.it (Stefano Bianchi)
Date: Thu, 15 Jun 1995 18:05:25 +0100 (MET)
Subject: Kurta Penmouse for Macintosh

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

Can anybody help a colleague of mine? The Kurta Penmouse for
Macintosh best run on the Mac with 2 Serial Port. This graphic tablet not
work on Mac Powerbook 150 because it have only one serial port?
This tablet required a specific driver to run on PW 150?

Many thanks in advance
Stefano Bianchi sbianchi-at-dbag.unifi.it
Dept. of Animal Biology
Univ. of Florence
Firenze, ITALY






From: Goldmarker-at-aol.com
Date: Thu, 15 Jun 1995 12:22:02 -0400
Subject: Coated grids for Acrylics?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Why is it necessary to use coated grids for acrylic embedding resins, but not
for epoxy resins?

What is the mechanism?

I would appreciate your help.

Thanks, Donald P. Cox




From: hunt-at-msc.cornell.edu
Date: Thu, 15 Jun 1995 13:06:58 -0400 (EDT)
Subject: silicon microtoming diamond knife

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Reply-To: hunt-at-msc.cornell.edu

Hi,

Does sombody know if one can cross-section a 25 micron

thick silicon piece with a diamond knife (microtoming).

Will that damage or dull the knife?

Thanks

jandt-at-msc.cornell.edu




From: hukee.margaret-at-mayo.EDU (Marge Hukee)
Date: Thu, 15 Jun 1995 13:46:39 +0200
Subject: Re: uranyl acetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear microscopists;
Just a note to update the uranyl crisis that we had encountered earlier.
Stacie Kirch was extremely helpful in clearing up the problem with uranyl
that was insoluble in water. We have tested a new batch supplied by both
EMS and Ted Pella and found that both suppliers now have uranyl that is
soluble at the 2% level. This is a vast improvement over previous lots of
uranyl.
Marge

Marge Hukee
EM Core Facility hukee.margaret-at-mayo.edu
Mayo Foundation (507) 284-3148
----------------------------------------------------------------------------
--






From: jester-at-crnjjsgi.swmed.edu (James V. Jester)
Date: Thu, 15 Jun 1995 11:09:44 -0500
Subject: Using same species AB

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sandy Masur last week had asked if any one knew how to do
co-localization with antibodies from the same species. I have
just come across an ad from Zymed for HistoMouse-SP which
supposedly contains a blocker for endogenous rodent IgG which
will enable one to use either rat or mouse IgG to localize
antigens. Perhaps a similar approach can be used to block
antigenic sites on the first antibody prior to application of
the second. has anyone used this HistoMouse-SP Kit. In the
long run I would think it would be preferable to make appropriate
antibodies in different species. I hope this might be of some
help.

_____________________________________________________________________
| | |
| James V. Jester | Dept Ophthalmology |
| jester-at-crnjjsgi.swmed.edu | UT Southwestern Medical Center |
| TEL (214)648-7215 | 5323 Harry Hines Blvd |
| FAX (214)648-2382 | Dallas, TX 75235-9057 |
|__________________________________|__________________________________|





From: Mr Herbert Mohr :      H.Mohr-at-ph.surrey.ac.uk
Date: Thu, 15 Jun 1995 18:50:44 +0100
Subject: SEM, EBIC, beam current measurements

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I do EBIC (Electron beam induced current) measurement on semiconductors, like GaAs,
InP, and other. One important parameter is the beam current in the micrcoscope. We
have an Oxford S250 with a normal tungsten cathode in our central microscopy facility.
The beam current is measured with a small Faraday cup and an external current amplifier.
The current is in the range of 0.1 to 5nA. But the beam is not always stable and I get
large fluctuations and/or drift in the current, which can be as big as 20%. I use the spot
size control, in order to get a certain beam current. But for the measurement on the
sample a fairly stable current is required.
Does anyone know, what one can expect as a good stable beam current and how to
achieve it? Or what might be wrong with the microscope that I experience such large
fluctuations.

cheers
Herbert Mohr
--------------------------------------
Science is the enemy of illusion.
-----------------------------------------------------------------------------
Herbert Mohr
Department of Physics, Optoelectronic and Devices group
University of Surrey
GU2 5XH
United Kingdom

e-mail: H.Mohr-at-ph.surrey.ac.uk
tel. : 01483/259403

SEM/EBIC - Monte Carlo simulation
------------------------------------------------------------------------------




From: cook-at-aaem.amc.anl.gov (Russell E. Cook)
Date: Thu, 15 Jun 1995 14:28:59 -0600
Subject: Re: old EDAX systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510102ac0648556a38-at-[146.139.72.78]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Concerning Scott Walk's recent comment (see below), EDAX has a program
called EDCVRT for its old PDP11 systems such as the PV9900: it translates
EDAX's EDS spectral files into ASCII format. Alan Sandborg of EDAX gave us
a copy 2-3 years ago. Also, the Fortran program NTRANS which Nestor
Zaluzec developed several years ago to translate various EDS spectral file
formats is still available from the anonymous ftp site WWW.AMC.ANL.GOV (IP
number 146.139.72.10). It may not be completely debugged: I had to
rewrite a small portion of it for the EDAX PV9900. I have taken Nestor's
code for the EDAX 9900 and modified it so that you are able to do batch
translations of EDAX's EDS files interactively. The ASCII files can then
be ported to a Mac or a PC over a serial line by a simple, albeit slow,
method. These text files have a MSA header and the data is in a single
column.

} Comment #3: With as many pieces of older EDS equipment that have been out
} there, the manufacturers should have written and distributed software to
} translate their spectra, data files, and image files to a standard format for
} both Mac and PC's. In fact, Nestor took the lead several years ago with
} establishing a format for EDS spectra, but no support from the manufacturers
} came about for older systems....
}
} JMHO
}
} Scott Walck

Russell E. Cook
Electron Microscopy Center for Materials Research
Argonne National Laboratory
9700 South Cass Avenue
MSD-212
Argonne, IL 60439
(708)252-7194
FAX: (708)252-4798






From: EMLAB-at-opus.mco.edu
Date: Thu, 15 Jun 1995 09:46:42 -0500 (EST)
Subject: TEM:PTA staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sorry about yesterday's lost message. Concerning the procedure about
negative staining of "polymere particles" with PTA. The procedure I have used
for viral particles is as follows: 1) Float a formvar-carbon coated grid on
drop of sample for 10-20 minutes. 2) Drain all excess fluid by touching edge
of grid to blotter paper. 3) Float grid on stain for 10 seconds and drain and
air dry.
PTA is a funny negative stain. It can be used at concentrations
ranging from 0.5-3% and at pH's between 4-8. You need to do various
combinations of the above and observe which is best for your sample.
You might want to try making a grid and not stain it. There might be
enough contrast so measurements can be obtained.

Hope this helps,

Ed Calomeni
Medical College of Ohio
Toledo, OH
emlab-at-opus.mco.edu






From: bergrh-at-cc.memphis.edu (R. Howard Berg)
Date: Thu, 15 Jun 1995 17:08:24 +0600
Subject: Re: manufacturers and old equipment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


WRT Scott Walck's comment 3 (below) that was in response to Don Grime's
message, I fully agree. I talked with the Tracor Northern people several
times in an effort to get help sending TN 5500 data to my Mac. They sent
me the file format for the FLEX images but made it understood that it was
up to me to find a programmer to get the files interpreted by the Mac.
They were interested selling us a newer Tracor system-but we've got more
pressing things to spend our budget on.

Howard Berg

Comment #3: With as many pieces of older EDS equipment that have been out
there, the manufacturers should have wirtten and distributed software to
translate their spectra, data files, and image files to a standard format for
both Mac and PC's. In fact, Nestor took the lead several years ago with
establishing a fromat for EDS spectra, but no support from the manufacturers
came about for older systems. This should have been done to include detailed
cabling and interface instructions. This should have been done with minimum
cost to the users who have to live with the systems that is some cases have
been purchased by others.

JMHO

Scott Walck


R. Howard Berg, Ph.D.
Biology Department
University of Memphis
Memphis, TN, 38152
E mail: bergrh-at-cc.memphis.edu
phone: 901-678-4449 fax: 901-678-4457







From: ingen-at-Rt66.com
Date: Fri, 16 Jun 1995 06:00:40 -0600
Subject: Re: manufacturers and old equipment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506161202.AA16896-at-Rt66.com}
X-Sender: ingen-at-rt66.com
X-Mailer: Windows Eudora Version 1.4.4
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

subscribe





From: SGKCCK-at-aol.com
Date: Thu, 15 Jun 1995 19:09:08 -0400
Subject: SILICON MICROTOMING WITH DK

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I saw your message regarding the above referenced subject. As Scott
suggested I do have some experiencing cutting this thickness of silicon.
However, it is rather lengthy and i do have some questions to ask before I
would make final recommendations. Please give me a call so we can
extensively discuss this.
I look forward to hearing from you.
Sincerely,
Stacie Kirsch
Diatome
Tel: 215-646-1478




From: anne-at-emu.su.oz.au (Anne Simpson Gomes)
Date: Fri, 16 Jun 1995 09:06:26 +1000
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


on holidays for a bit, please unsubscribe

thanx

Anne

Anne Simpson Gomes

EM Unit, F09
Univ of Sydney | OVE
NSW 2006 Australia lots of | UCK and
Fax: (612) 552 1967 |____ AUGHTER ASG







From: Diane Montpetit :      MontpetitD-at-EM.AGR.CA
Date: Fri, 16 Jun 1995 11:25:52 -0400
Subject: mocha image analysis overlapping globules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {sfe16ae6.073-at-EM.AGR.CA}
X-Mailer: Novell GroupWise 4.1

hello everyone,

I am currently working with an image analysis system called MOCHA,

I am looking at light microscopy pictures of fat globules of varying sizes and trying to
measure their average diameter, but some of them are overlapping and I just can find
out how to get the image analysis system to take this into account...The boundary is
visible...Their is an exemple of coins touching to each other but nothing concerning
transparent objects overlapping.

I am wondering if the system is able to do what i want him to do...



Diane Montpetit
email;montpetitd-at-em.agr.ca
fax 514 773 8461
tel514 773 1105

food research center
quebec, canada






From: Matt Kizerian :      kizerian-at-ucsu.Colorado.EDU
Date: Fri, 16 Jun 1995 09:26:32 -0600 (MDT)
Subject: Where is Bio Rad/Polaron?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In the Chemical Engineering Dept. of the University of Colorado at
Boulder we are trying to use our Bio Rad/Polaron E7400 cryostage system on
our Cambridge SEM.

Unfortunately it has not been used in a while & we need to reinstall part
of it & need documentation on its use. The PROBLEM is that we can't reach
Bio Rad ANYWHERE to request manuals. We've tried the following phone
numbers, but haven't even reached a recording at any of them:

617-864-5809
617-864-5820
1-800-524-8200

Is Bio Rad still out there? If so, where can they be reached? We're trying
to get some important lab protocols worked out & need to get this
equipment up and running ASAP. Thanks for any help!


Matt Kizerian
NSF Center for Seperations Using Thin-Films
Chemical Engineering Dept.
University of Colorado at Boulder




From: Sally Shrom :      sally-at-retina.anatomy.upenn.edu
Date: Fri, 16 Jun 1995 13:22:02 -0400 (EDT)
Subject: TEM / Bio / Uranyl Acetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Posted-Date: Fri, 16 Jun 1995 13:23:29 -0400

Uranyl Acetate: how hot is it?




From: Greg :      GREG-at-umic.umic.sunysb.edu
Date: Fri, 16 Jun 1995 13:26:05 EST5DST
Subject: IEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a protocol for the fixing of bone for gold labelling of
osteocytes in undecalcified bone. I am going to be doing TEM.

Greg Rudomen
University Microscopy Imaging Center
University at Stony Brook, New York
Greg-at-umic.umic.sunysb.edu





From: Greg :      GREG-at-umic.umic.sunysb.edu
Date: Fri, 16 Jun 1995 13:46:07 EST5DST
Subject: mocha

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think you will have to crop out each globule. Then paste them into
one image and then do your measurements. I don't believe there is a
program to do what you want automaticly.

Greg Rudomen
UMIC
University at Stony Brook
Greg-at-umic.umic.sunysb.edu





From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 16 Jun 1995 14:39:33 -0400
Subject: RE- Bio-Rad Addr

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1408808702.55522-at-mse.engin.umich.edu}

Subject: Time: 2:32 PM
OFFICE MEMO RE: Bio-Rad Addr Date: 6/16/95

The 1995 list of manufacturers published by R & D Magazine has two entries
for Bio-Rad:
Digilab Div, 237 Putnam Ave., Cambridge, MA 02139.
Ph: 617-868-4330, and
Sadtler Div, 3316 Spring Garden St.,Philadelphia, PA 19104
Ph:215-382-7800
Good Luck!





From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Fri, 16 Jun 1995 09:17:43 -0500
Subject: SPM (any type)/plant cell walls

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
Is there anyone out there who is doing (or who knows of someone
doing) spm on isolated plant cell walls? I am interested in this myself and
I would like to hear from anyone who has had experience with this type of
sample.

Thanks in advance,
Tobias Baskin

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
___ ____ ^ ____ _____ Tobias I. Baskin
/ \ / / \ / \ / University of Missouri
/ | / / \ / / Biological Sciences
/___ / /__ /_____\ / /__ 109 Tucker Hall
/ / / \ ( / Columbia, MO 65211 USA
/ / / \ \ / voice: 314-882-0173
/ /____ / \ \____/ /_____ fax: 314-882-0123






From: jfmjfm-at-umich.edu (John F. Mansfield)
Date: Fri, 16 Jun 1995 09:54:50 -0400
Subject: TN/Noran5500 to Mac

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This ought to be in the FAQ list!

In fact this applies to any PDP-11 based XEDS system.
In the PDP-11 put an ethernet card in the QBUS. Get a license for RT-11,
not difficult since people are tossing out PDP-11 systems left and right.
Buy a copy of ftp for RT-11 from Process Software. Run ftpd on the RT system
Run a strip of ethernet from the PDP-11 to the Mac (or PC or Unix box) and
run Fetch (or your favourite ftp client) and list the files on your pdp
hard disk and then select the ones you want and pull them over. Works very
will and you can do images too!

Contact me of line for further information.

OK?

John Mansfield
North Campus Electron Microbeam Analysis Laboratory
413 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX (313)936-3352
jfmjfm-at-engin.umich.edu, John.F.Mansfield-at-umich.edu
or jfmjfm-at-umich.edu they all reach me!
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html






From: JOHNA-at-SCI.WFEB.EDU
Date: Fri, 16 Jun 1995 14:42:47 -0400 (EDT)
Subject: Re: Where is Bio Rad/Polaron?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Try calling Energy Beam Sciences (800 992-9037; fax 413 789-2786). I
believe that they took over at least part of Bio-Rad/Polaron's line of
ancillary instrumentation.

Good Luck

John

___________________________________________________
| |
| John G. Aghajanian, Ph.D. |
| Worcester Foundation for Experimental Biology |
| 222 Maple Avenue |
| Shrewsbury, MA 01545-2737 |
| |
| Tel: 508 842-8921 ext. 147, 161 |
| Fax: 598 842-9632 |
| JOHNA-at-sci.wfeb.edu |
| |
|_________________________________________________|





From: wise-at-vaxa.cis.uwosh.edu
Date: Fri, 16 Jun 1995 16:04:07 +0000
Subject: EDS of bone medullary tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To all,

We are preparing to use EDS to look for lead in the medullary bone
tissue of geese poisoned by lead shot ingestion. The medullary tissue is
so fragile that we will probably have to embed whole bones in epoxy and
then slice the bone in half. Does anyone have any advice on which epoxy is
preferred? Would polishing of the cut face be necessary? What is used for
a polishing compound? Do the polishing compounds leave contamination?
Etc?
Any help would be appreciated.

Bob Wise
Department of Biology
University of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404
(414) 424-1101 fax
wise-at-vaxa.cis.uwosh.edu






From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Fri, 16 Jun 1995 16:39:35 -0600
Subject: SEM:using virus as surface label

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v01510101ac07b7f9276b-at-[131.230.97.56]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We are revising a textbook and are looking for a SEM micrograph
illustrating the use of viral particles (or possibly haemocyanin) for
labeling cell surfaces. Can anyone help? Thank you.

John J. Bozzola
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
Phone: 618-453-3730
Fax: -2665
Email: bozzola-at-siu.edu OR bozzola-at-qm.c-cem.siu.edu






From: William Tivol :      tivol-at-wadsworth.org
Date: Fri, 16 Jun 1995 16:23:36 -0500 (EDT)
Subject: Re: How hot is Uranyl acetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sally,
My Radiation Sciences class does something like this every year.

1) Look up the half-lives and fractions of uranium isotopes (ignore
14C, 3H and 15O, etc.).
2) Calculate the decay rates from lambda = .693/halflife.
3) Calculate the number of uranium atoms in a unit volume of solu-
tion (this depends, of course, on the strength of the solution).
4) Calculate the number of decays from dN/dt = - lambda N.
5) Sum the contributions from each isotope (if necessary, or if
anal retentive).
6) Convert to Ci/l (or other familiar units).

We treat small amounts of 1% uranyl acetate as low-level waste,
which can be poured down the sink. The short answer, therefore, is
"Not very hot."
Yours,
Bill Tivol




From: m.dickson-at-unsw.edu.au (Melvyn R. Dickson)
Date: Fri, 16 Jun 1995 10:38:44
Subject: Re: Coated grids for Acrylics?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In article Goldmarker-at-aol.com writes:


} Why is it necessary to use coated grids for acrylic embedding resins, but not
} for epoxy resins?

The electron beam breaks chemical bonds and heats the specimen locally to
around 200 deg celsius. Even epoxy resins lose about half their mass by
evaporation under irradiation before they stabilise as a non volatile
residue. This explains why thicker sections tend to "clear" with the
background becoming more transparent and the metallic stain precipitate
becoming denser (it is becoming physically denser) as the beam hits them.
Acrylic resins are much less stable in the electron beam than epoxy resins
and melt and lose mass to the extent they lodse their mechanical strength and
tear and shrink up to the grid bars.

This ties in with the phenomenon of contamination, in which the vaporised
organic materialsfrom the section redeposit on the section, where they are
depolymerised into less volatile carbon by the beam. This adds mechanical
stability at the cost of some loss in contrast. You can stabilise your
sections this way, but its more controlled to put a very thin carbon coat on
your sections (maybe both sides) in a carbon coater.




From: GVKM07A-at-prodigy.com (DR CHARLES A GARBER)
Date: Thu, 15 Jun 1995 23:42:30 EDT
Subject: TEM, Si thin sectioning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

On June 15, it was asked "if one can cross-section a 25 micron thick
silicon piece with a diamond knife (microtoming)". And it was also
asked "Will that damage or dull the knife?"

1] One can do it, however it takes a lot of practice, and even then
the quality of the sections is always less than what would be desired
(the section tends to break up because it is so brittle, yet not that
badly that one can not get useful information)

2] It definitely will "damage" or "dull" the knife. You should
collect damaged diamond knives from colleagues and use them for the
rougher part of the operation, e.g. "facing off" the block, for
example. I could tell you horror stories where the user "broke" the
diamond in the knife on his first attempt. Yet others have related the
opposite experience, that is, they were able to cut some number of
samples before the knife was totally kaput.


We have been using our own SPI Materials Science diamond knife for this
kind of sectioning, but I would expect that other "brands" of diamond
knives could be used as well. If you sent us a small piece we could
determine to what degree one could actually do this with your
particular kind of sample in our own "working" laboratory and send you
back a few grids for you to "inspect" in your own TEM. If we could cut
your type of sample satisfactorily, then I would expect that you could
eventually do it as well.

======================================================
Charles A. Garber, Ph. D.
President
SPI SUPPLIES
PO BOX 656
West Chester, PA 19381-0656 USA

Ph: 1-(610)-436-5400
1-(800)-2424-SPI

FAX: 1-(610)-436-5755

e-mail: GVKM07A-at-prodigy.com
SpiSupp-at-aol.com [SPI Customer Service e-mail box]
======================================================







From: A. Kent Christensen :      akc-at-umich.edu
Date: Thu, 15 Jun 1995 15:26:21 -0400 (EDT)
Subject: Re: EM AND LM IMMUNOSTAINING: Ab to FITC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jay,

You might want to check with Amersham (technical 1-800-341-7543,
order 1-800-323-9750). They have a non-radioactive ISH system based on
using fluorescein as the marker on the nucleotide, then ICC detection of
the fluorescein with a Fab fragment labeled with alkaline phosphatase
(BCIP/NBT). Your ICC would be equivalent to the detection step in their
system. I don't know about EM (ask them).

A. Kent Christensen, University of Michigan, {akc-at-umich.edu}

--------------------------------

On Thu, 15 Jun 1995, Jay Jerome wrote:

}
} I have a client who wishes to immunostain for FITC. He would like to do
} this at the light microscope level in both parafin and fixed,
} cryosectioned material. He would also like to do this at the EM level by
} pre-embed immunostaining.
}
} My questions are:
}
} 1. Any recommendation on which of the Anti-FITC anti-bodies work well (or
} don't work) for immunhistochemistry? [species the antibody was raised in
} is not limiting at this point]
} a. Parafin embedded material?
} b. Crosections?
} c. Fixed, but unembed material?
}
} 2. Can anyone provide a good reference (I have a couple of papers but the
} methods are sketchy) or protocol for any of the three above staining
} situations with Anti-FITC? Concentrations, times, etc?
}
} 3. Are there any unusual pitfalls to worry about? We have considerable
} experience with immunostaining, so have the basics down.
}
} I appreciate any help y'all can provide. Thanks
}
} Jay Jerome
} **************************************************************
} * aka: W. Gray Jerome *
} * Dept. of Pathology *
} * Bowman Gray School of Medicine of Wake Forest University *
} * Medical Center Blvd *
} * Winston-Salem, NC 27157-1092 *
} * 910-716-4972 *
} * jjerome-at-isnet.is.wfu.edu *
} **************************************************************
}
}




From: dlb-at-aruba.ccit.arizona.edu (David Bentley)
Date: Fri, 16 Jun 1995 17:03:37 -0700
Subject: Re: TEM / Bio / Uranyl Acetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Mallinkrodt MSDS sheet lists the specific activity of UA as
around 0.2 microcuries/gm. Meter readings from our Geiger counter show 300
cpm (.1 mrem/hr) about 2 cm from the side of a 5gm bottle and 8,000 cpm (~3
mrem/hr) 2 cm
from the top of an open bottle. The glass shields the weak x-Ray (0.048
mev) fairly well. A quarter pound bottle had 1,000 cpm (~.4 mrem/hr) at the
sides and 20,000 cpm (~10 mrem/hr) at the open top. Larger quanities of UA
would have to be scaled up appropiately.

In noting another response, you may want to check the local
ordanances, is is possible that the sewer stream in your area is restricted.
Unless my quick calculations are wrong (and they often are) even 1 drop of
saturated UA would exceed our restrictions of 12 picocuries/day. I would
agree that compared to most other isotopes, UA is a low level radiation
source (but it's hazards, especially as a heavy metal, should not be ignored)

Off the topic but for the question of UA not going into solution
asked earlier, one solution is given in Millonig's book Laboratory Manual of
Biological Electron Microscopy, (1976). Adding a drop or two of glacial
acetic acid per 100 mls of water will help the covert the insoluable forms
to soluable UA. It also helps prevent precipitation and increase shelf
life. We have a stock bottle of saturated UA solution in a 1 L brown
bottle, which periodically is added more DI water and more UA. From this is
decanted what is needed and lower % solutions diluted. UA is saturated at
7.7% (15 C.)(Hayat 3rd ed), for room temp, ~8%.




later
dlb





From: GVKM07A-at-prodigy.com (DR CHARLES A GARBER)
Date: Fri, 16 Jun 1995 23:51:36 EDT
Subject: Bio-Rad/Polaron?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

On June 16, Matt Kizerian asked what happened to "Bio-Rad/Polaron"?

Several years ago, the EM lab "prep" business, which was part of the
Microscience Division of Bio-Rad was sold to Fisons Instruments, and
once over at Fisons, they divided the business into three parts:

a) Histology equipment which was sold to Energy Beam Sciences

b) EM equipment business which was taken over by Fisons VG Microtech,
and their distributor in the USA is Energy Beam Sciences, and

c) Polaron consumables which was taken over by Fisons "corporate". In
Sept. 1994, the entire inventory of the consumables business was
purchased by SPI Supplies, and is being offered to those who want to
continue to use some of the unique and different items from the Bio-Rad
EM consumables catalog not available anywhere else.


You can reach Energy Beam Sciences on (800) 992-9037. They should be
able to help you with what you need. If not, let me know and I will
try to get what you need. In our own laboratory we have an
Oxford/Hexland system and I would imagine the "protocols" are virtually
the same.

For your information, Bio-Rad is very much "still out there" however
the part that we had been all dealing with as part of their "EM lab
prep business" has not been a part of Bio-Rad for some several years.

======================================================
Charles A. Garber
President
SPI SUPPLIES
PO BOX 656
West Chester, PA 19381-0656 USA

Ph: 1-(610)-436-5400
1-(800)-2424-SPI

FAX: 1-(610)-436-5755

e-mail: GVKM07A-at-prodigy.com
SpiSupp-at-aol.com [SPI Customer Service e-mail box]
======================================================







From: Zhiyu Wang :      zhiyu-at-uhunix.uhcc.Hawaii.Edu
Date: Sat, 17 Jun 1995 01:19:53 -1000
Subject: Re: mocha image analysis overlapping globules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

This message is for Mr. Diane regarding the usage of MOCHA. I have
experience in measurement of images using MOCHA. It is hard to have
the program to identify the object being overlayed, but you can "help" it.
Go to annotation function, choose free draw, select one of four overlay
color, and draw on the object bandary, then fill out with the overlay color.
Now you can get measurement automaticaly by measuring the overlay object.
You can draw two objects being overlayed by using different color and
signe the colors as two source to be measured.

Good luck!

Wang




From: gt0194g-at-prism.gatech.edu (Joseph E. Coury)
Date: Sat, 17 Jun 1995 10:09:44 -0400
Subject: SPM WWW SITES

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am currently undertaking the task of compiling a list of World Wide Web
sites dealing with SPM and "Nano" technologies. Is there anyone out there
who has a similar list that might be willing to share it with me? Thank-you
in advance for any help that you can give me.
---------------------------------------------------
Joseph E. Coury
School of Chemistry and Biochemistry
Georgia Institute of Technology
Atlanta, GA 30332-0400 USA

(e-mail) gt0194g-at-prism.gatech.edu
(FAX) 404-894-7452
(LAB) 404-894-4064
---------------------------------------------------





From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Sat, 17 Jun 1995 11:57:16 -0400 (EDT)
Subject: RE: SPM (any type)/plant cell walls

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Fri, 16 Jun 1995 09:17:43 -0500,
baskin-at-biosci.mbp.missouri.edu (Tobias Baskin) writes:

} Greetings,
} Is there anyone out there who is doing (or who knows of someone
} doing) spm on isolated plant cell walls? I am interested in this myself
___________

Please contact Dr. Thomas Pesacreta, Microscopy Center, University of South
Western Louisiana, PO Box 42451, Lafayette, LA 750504-2451 (Tel: 318 231-5769;
Fax: 318 231-5864). I believe he has or he is currently studying plant cell
walls using an SPM.

*************************************************************************
M.V. Parthasarathy
Professor of Plant Biology & Director, EM Facility
Section of Plant Biology
228 Plant Science Building
Cornell University, Ithaca, NY 14850
Telephone: (607) 255-1734
Fax: (607) 255-5407
E-mail: mvp2-at-cornell.edu
***********************************************************************




From: Zhiyu Wang :      zhiyu-at-uhunix.uhcc.Hawaii.Edu
Date: Sat, 17 Jun 1995 20:28:13 -1000
Subject: Re: mocha image analysis overlapping globulesp

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I am responding to Diane Montpetit message regarding to the usage of

MOCHA softwaer. I have some experience in setting and using of this

softwaer. It is hard to have the prograsm to identify the objects

overlayed, but you can "help" to do it. Go to ANNOTATION function and

select free draw tool. Draw on the bandary of the object then fill it

with one of four color. Draw another object and fill it with different

color. Now you can have the program to measure the object in different

overlay of color by setting the colors as a source in MEASUREMENT SETTING

window.

Let me know if you have better idea. Good luck!

Wang




From: Dr. Molnar Peter :      MOLNARP-at-lib.dote.hu
Date: Mon, 19 Jun 1995 10:14:16 +100
Subject: Fe identification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Folks:
Is there anyone out there who knows a specific method that is usable
in conventional TEM for unequivocal identification of iron? I have a
case of chronic granulomatous disease with Prussian blue deposition
in macrophages. I would like to have a similar reaction done in EM.
Any suggestions would be welcome.
Thanks
Peter Molnar M.D., Ph.D.
Hungarian-Japanese EM Ctr
Dept. Pathology, Univ. Med. Sch.
Debrecen, POB. 24.
Nagyerdei krt. 98.
H-4012 Hungary, Europe
FAX/Tel:36-52-417-063
e-mail:molnarp-at-lib.dote.hu




From: randio-at-fagmed.uit.no (Randi Olsen)
Date: Mon, 19 Jun 1995 15:13:49
Subject: EM AND LM IMMUNOSTAINING: ab to FITC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Last week, Jay Jerome had some questiones about antibodies to FITC.

We have been working with endocytosis of circulating collagen; labelling
collagen with FITC and injection of this collagen into a fish.
For LM detection of the FITC/collagen we just fixed the tissue, embedden in
parafin, sectioned and put it under a fluorescense microscope.
If you want to stain the tissue, make sure you don't use eosin, its a
fluorochrome itself.
For EM studies we used cryosections (fixed with 8% formaldehyd and 0,5 %
glutaradelhyde) Monoclonal Anti FITC from Boehringer Mannheim Biochem,
Germany worked well at a dilution of 1: 100. Bridging antibody not needed.
We used a standard Protein A-gold prosedure.

We also did a adsoption control, and got no signal.

For more details, see Cell and Tissue Research (1995) 280:39-48 ; Smedsrod
et al: Circulating collagen is catabolized by endocytosis mainly by
endothelial cells of endocardium in cod.

Best regards
Randi Olsen
University of Tromso
Department of Electron Microscopy
MH-Breivika
N-9037 TROMSo
Norway





From: TIETZ :      100115.66-at-compuserve.com
Date: 19 Jun 95 09:17:34 EDT
Subject: TEM Cryo Workshop 1995

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


All interested microscopists are invited to the CRYO-MICROSCOPY WORKSHOP
from August 28th up to September 1st 1995 organised by the Dutch Society of
Electron Microscopy and Philips Electron Optics

In view of the increasing interest and in order to further cryo-electron
microscopy,
special workshops have been organized since 1989 every two years by Philips
Electron
Optics in cooperation with the Electron Microscopy Society of The Netherlands.
The next workshop will be from August 28th up to September 1st. Experts in the
various fields will teach many aspects of cryo electron microscopy in
theoretical
and perhaps more importantly practical sessions. The workshop will be held at
the
Philips Electron Optics Applications Laboratory with its wide range of essential

equipment, including microscopes, ancillary equipment. Other, specialized
equipment
is made available by several other manufacturers.
A number of cryo-experts is invited as teachers for the workshop.

TOPICS

Cryo-ultra microscopy and Immunogold labelling Freeze substitution.
Cryo-electron microscopy:
- Image Analysis of Spherical Virus.
- Automatic electron tomography of ice-embedded molecules.
- Energy filtering of isolated biological macromolecules.
- 2 Dimensional Protein Crystals in Ice.
- Electron Dosis and the use of a Slow Scan C.C.D.

INVITED SPEAKERS

A. Brisson, T. Baker, B. Koster, H. Gross, W. Voorhout, J. Leunissen, B. Humbel,
G. Knoll, J.W. Slot.

ORGANISERS

Prof. Dr. A. Verkleij, University of Utrecht
Dr. P. Frederik, University of Limburg
W. Busing, Philips Electron Optics

FORMAT

The workshop duration will be 5 days. For the first 4 days, each activity will
include two one hour sessions as an introduction for the practical sessions.
The practical sessions will be organised by cryo-specialists from the
Universities of Utrecht and Limburg and participants will have the opportunity
for hands-on-practise and to work with their own specimen.

The fifth day, the morning will also start with theory sessions and after this,
the cryo-course will be evaluated. In the afternoon the program is open to give
the participants time for travel arrangements and/or to continue to work with
their own specimen.

Registration form available from (the dead line is July 1,1995):

W.M. Busing
Philips Electron Optics
Building AAE
P.O. Box 218
5600 MD Eindhoven
The Netherlands
Fax: +31 40 766102 or

Tietz Video & Image Processing GmbH
Herbststr. 7
D-82131 Gauting
Phone: 089/8506567
Fax: 089/8509488

e-mail: 100115.66-at-Compuserve.COM






From: A. Kent Christensen :      akc-at-umich.edu
Date: Mon, 19 Jun 1995 09:56:34 -0400 (EDT)
Subject: Re: Using X-gal for ?-galactosidase demonstration at EM level

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Terry,

I have heard that time in propylene oxide needs to be reduced. Some
references that include EM of X-gal product:
Engelhardt et al., 1991, PNAS 88:11192.
Franklin & Barnett 1991, Acta Neuropath 81:686.
Loewy et al. 1991, Brain Res 555:346.

A. Kent Christensen, University of Michigan, {akc-at-umich.edu}

-------------------------------------------

On Fri, 16 Jun 1995, T. Robertson wrote:

}
} To any myologists out in cyber space
} I am attempting to demonstrate ?-galactosidase in a strain of lacZ mice
} (transgenic mice strain for fast troponin) using X-gal histochemistry and
} conventional processing for transmission electron microscopy. Although I can
} clearly see labelled nuclei at the light microscope level I am having
} trouble routinely seeing positive cells at the TEM level. Does anyone have
} experience in this field and is there a step in the processing that might
} remove much of the reaction product. We dehydrate in graded ethanol, use
} propylene oxide as a transitional solvent and embedd in Araldite. I would
} be grateful for any helpful comments.
}
} Terry
}
} Dr Terry Robertson
} Electron Microscopist
} Department of Pathology
} University of Western Australia
} Nedlands 6009
}
} phone 346 2935
} Fax 346 2891
} email troberts-at-eosin.path.uwa.edu.au
}




From: cook-at-aaem.amc.anl.gov (Russell E. Cook)
Date: Mon, 19 Jun 1995 11:10:03 -0600
Subject: electropolishing cast iron

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510101ac0b5e65bbc8-at-[146.139.72.78]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

1. See Ultramicdoscopy 25 (1988) 89-90. It describes 65 ml HCl, 435 ml
methanol, and 20 ml butyl cellosolve (a viscosity improver) at -20 C, 110
volts, 120 mA on a single jet South Bay instrument.

2. Also try: 50 ml perchloric acid, 450 ml acetic acid, and 50 ml butyl
cellosolve at room temp., 42 volts, 24 mA. Again, this was on a South Bay
550-B single jet instrument. Great results on cast stainless steel,
furnace aged.

Who has a good D.C. electropolish for platinum ??? A.C. in the literature!!


}
} *From: "A. HONARBAKHSH-RAOUF" {CER5AH-at-ECU-01.NOVELL.LEEDS.AC.UK}
} *To: microscopy-at-aaem.amc.anl.gov
} *Subject: cast iron thin foil
}
} *Does anyone know any suitable solution for preparation of thin
} *foil using jet polishing of ductile cast iron(presence of two
} *different phases ie free graphite and matrix)? Is there any other
} *method except ion beam?
} *Thanks
} *Abbas

Russell E. Cook
Electron Microscopy Center for Materials Research
Argonne National Laboratory
9700 South Cass Avenue
MSD-212
Argonne, IL 60439
(708)252-7194
FAX: (708)252-4798






From: John.Wheatley-at-ASU.Edu (John C. Wheatley)
Date: Mon, 19 Jun 1995 09:40:59 +0800
Subject: Legal/Ethical Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: wheatley-at-csss.la.asu.edu
Message-Id: {v01510100ac0a8599d325-at-[129.219.44.209]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I am interested in the legal and ethical arguments (pro and con) for
university involvement with industrial entities in the area of electron
microscopy. In other words, is it legal/ethical for universities to use
university instruments to solve problems for industry on a fee-for-service
basis? I would like to have replies from anyone who wishes to respond.
However, I realize that regulations may be different in other countries and
therefore I ask that your responses be sent directly to me so our
international friends are not bored with American legal/ethical
discussions. I will provide a compilation of the responses to all who ask
for them, except when individual respondents request otherwise.

With respect to the legal side of the question, I would like to have
answers based upon actual legal opinions that you are aware of.
Perceptions about legality will not be useful. I know that there are
probably very few (maybe none) lawyers involved in electron microscopy. If
your institution has issued a legal opinion with respect to this issue, I
would like to know what that opinion is.

I would find it useful to have many people respond to this request. I
would especially like to have those members of the private sector, who use
electron microscopes to solve problems for industry, respond.

If you feel comfortable responding, please include your price structure for
industrial jobs--proprietary and otherwise.

I thank you for your responses.

John C. Wheatley
Lab Manager
Center for High Resolution Electron Microscopy--Arizona State University

John C. Wheatley
Lab Manager
Center for High Resolution Electron Microscopy
BOX 871704
Tempe, AZ 85287-1704
Phone: (602) 965-3831
FAX: (602) 965-9004
John.Wheatley-at-ASU.Edu






From: Jaime Dant :      jaime-at-borcim.wustl.edu
Date: Mon, 19 Jun 1995 12:55:58 -0500
Subject: re: how hot is ua

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Good afternoon Sally,

I purchase UA from EM Sciences and they print right on the bottle its activity
as 0.51 uCi/gm.

Jaime A. Dant




From: wzp-at-befvax.uchicago.edu
Date: Mon, 19 Jun 1995 12:59:19 EDT
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Zhiping Wang from the University of Chicago





From: modb-at-ruca.ua.ac.be (Marc Op de Beeck)
Date: Mon, 19 Jun 1995 21:00:40 +0200
Subject: HREM SIMULATIONS COMPARISON

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear HREM collegues,

Together with Dr. Dirk Van Dyck, I started a comparison of different
HREM diffraction simulation programs in order to check consistency.
After some "false" starts (personal mailings, request in Ultra-
microscopy, ...) on which we got hardly no reaction at all, we finally
decided to start the comparison with the programs we had available at
the Universities of Antwerp (Belgium) and Amsterdam (The Netherlands).

Together with collegues in Amsterdam/Delft University, we tested some
commercially available programs we had access to, such as :

* MacTempas : done
* Cerius2 : done
* EMS Multislice : in progress
* EMS Bloch Wave : in progress

against some non-commercial ones, as :

* Real Space : done Dirk Van Dyck, Wim Coene, ...
* ONERA, Bloch wave : done Cyrille Barreteau
* ONERA, Multislice : done Cyrille Barreteau
* TCBED : done Zuo and Spence
* Dong, Multislice : done Dong Tang
* Chen, Multislice : in progress Jianghua Chen
* NCEMSS Multislice : in progress Roar Kilaas

With these results, I created a WWW site "http://www.ruca.ua.ac.be/~modb"
which has been made known to the commercial authors of the following
packages :

* Kilaas : MacTempas, NCEMSS
* O'Keefe : SHRLI
* Molecular Simulations : CERIUS2
* Stadelmann : EMS
* Epicier : SIMPLY (shareware)

Seeing these results, all authors finally have agreed to contribute to
the project (and if necessary, change their programs according to the
requirements of the proposal such as the use of different potentials).
Up till now, the comparison with MacTempas and the CERIUS2-HRTEM packages
have been finished in collaboration with the respective authors. EMS will
be included shortly (in he next two weeks) in the overview. The present
results show that all packages can reproduce the same results, provided
they are properly used (which quite often means that using the standard
values in these programs is NOT appropriate to get fully converged
results).

It is however clear that not all currecntly used packages have been
included in this survey. We simply choose for the packages above, since
they were available to us. Therefore, all authors and users of these
"neglected" packages are kindly requested to contribute to this
comparison. The full text of the proposal can be found in
* "Ultramicroscopy 55 (1994), 435-437"
* WWW site "http://www.ruca.ua.ac.be/~modb"


Sincerely yours,

Dr. Marc Op de Beeck
University of Antwerp (RUCA-EMAT)
Groenenborgerlaan 171
B-2020 Antwerpen
BELGIUM

Tel : ++ 32 3 218 02 61
Fax : ++ 32 3 218 02 57
URL : http://www.ruca.ua.ac.be/~modb
E-mail : modb-at-ruca.ua.ac.be






From: c4647-at-slvaxa.umsl.edu (Phil Fraundorf)
Date: Mon, 19 Jun 1995 14:15:28 -0500
Subject: Re: Legal/Ethical Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Yo John,

You wrote for the ANL microscopy listserver:
} I am interested in the legal and ethical arguments (pro and con) for
} university involvement with industrial entities in the area of electron
} microscopy. In other words, is it legal/ethical for universities to use
} university instruments to solve problems for industry on a fee-for-service
} basis? I would like to have replies from anyone who wishes to respond.
} However, I realize that regulations may be different in other countries and
} therefore I ask that your responses be sent directly to me so our
} international friends are not bored with American legal/ethical
} discussions. I will provide a compilation of the responses to all who ask
} for them, except when individual respondents request otherwise.

It's NICE that you should ask about ethics, even if legal had a tendency
to come first. I have some experience with mandates on "both sides of the
aisle", and have in fact recently discovered that development of a "Code of
Ethics for Analytical Support" can have important PRACTICAL consequences for
the kind of collaborations you describe, as well as for the development of
constructive legal precedents in days ahead. Look for benefits to
university, industry, and even instrument researchers, as well as to the
appropriate pooling of resources between institutions.

Using as template codes developed in other fields (e.g. occupational
therapy, for example), we are in the process of preparing an article with
some specifics. One objective of this initiative is to provide a mechanism
which will allow us to avoid asking questions of lawyers BEFORE they are
equipped to answer them (something that generally has a negative impact on
everyone involved). In that context I expect that what you will find out
there now, in your survey, may not be terribly enlightened. Nonetheless, I
am interested, and will try to keep you posted on our initiatives in this
vein as well.

One question I have of LISTSERVER MEMBERS in this context is: What work
on a code of ethics for researchers in analytical support (like microscopy)
is already available to draw from? (As with John's request, feel free to
reply directly rather than through the listserver.)

Cheers. /philf :)

//\/\/\/\--}
// Phil Fraundorf, Physics & Astronomy: 3145165044 c4647-at-slvaxa.umsl.edu
\\ U. Missouri - St. Louis MO 63121 USA http://newton.umsl.edu/pfhomepg
\\/\/\/\/\/\/\/--}





From: wise-at-vaxa.cis.uwosh.edu
Date: Mon, 19 Jun 1995 15:37:17 +0000
Subject: Legal/ethical question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to John.Wheatley's Legal/Ethical Question:

The legislature of the State of Wisconsin has addressed this issue
and has written statutes that very clearly delineate what university
facilities can and cannot be used for and under what terms. Basically,
non-university users can rent beam time but they must be charged a
competitive rate so as not to undercut private testing labs. I have
responded in much more detail directly to John.

Another side of the issue, however, is the direct impact on the
laboratory director. I am an assistant (tenure-track) professor at the
Uniersity of Wisconsin Oshkosh. I am allowed to do as much consulting as I
wish and charge whatever rate I want to as long as it does not interfere
with my job duties as described in my contract and the Faculty Handbook.
The Handbook says that I will be evaluated (for retention, promotion,
tenure, merit pay increases, etc) 45% on my teaching, 45% on research and
10% on service to the governance of UWO. Performing research for private
industry is not included anywhere in my job evaluation so even though I may
make a zillion dollars an hour, I am doing nothing to gain tenure. I
suppose that if I were Professor Super Microscopist I could handle all of
my teaching duties, publish 10 times a year, head the Faculty Senate, and
pull in $200,000 a year in outside consulting. But I'm not.

In short, although my dean has not forbidden me from taking on
outside work, he has made it very clear that I will get no credit for it
when it comes tenure time. I have ocassionally done this sort of work for
people or institutions (most of it gratis) that really needed my help but,
on the whole, I avoid it if I can.

Bob Wise
Director, UWO EM Facility
Univresity of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404
(414) 424-1101 fax
wise-at-vaxa.cis.uwosh.edu






From: mager-at-unixg.ubc.ca (Mary Mager)
Date: Mon, 19 Jun 1995 23:26:08 -0800
Subject: legal/ethical question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: mager-at-pop.unixg.ubc.ca (Unverified)
Message-Id: {v01510104ac0c2040beda-at-[137.82.220.138]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear John,
I run a multi-user Materials EM lab at the U. of British Columbia and this
question has come up in the past, particularly when the first commercial
Failure Analysis consultant bought his own SEM and hired my predecessor
away. The deal worked out with this consultant and consolidated since is:
1. The undergrad labs have first priority in the university lab, the
graduate student work the next. The other university users, who pay an
hourly fee ($25/hr) far less than the commercial rate, have next priority.
2. Commercial work, when done in time extra to that needed for university
work, is charged at a rate exceeding that of the local commercial
suppliers. It is important that there is no unfair competition with
government-funded equipment.
There were no legal precedents that I could find at the time this
arrangement was set up (1981). The consulting company sometimes finds it is
necessary to have another lab perform tests on a sample if they represent
one side of a litigation, so our relationship with the "competing" lab is
friendly. I send work to them if it is more their thing, they send them to
me if it is something I can do better.
The service I provide is strictly beam time, photos, EDX analysis and a
brief note on what I found. I am a technician, not a professor, so this is
not a consulting job, rather a lab service.
I have heard the opinion from some Americans that this kind of arrangement
is considered criminal conflict of interest, but here there doesn't seem to
be a problem. You were not very clear on the exact nature of the
relationship you wish to establish with the industrial entities, but I
cannot see the benefit in forbidding it altogether. The money coming in,
since it is at a high commercial rate ($200/hr.), is very welcome to help
run the lab.
I hope this overly-long reply is some help.

Mary Mager
Electron Microscopist
Metals and Materials Eng, UBC
309 - 6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648, fax: 604-822-3619
email: mager-at-unixg.ubc.ca






From: Mike Witcomb :      MIKEW-at-gecko.biol.wits.ac.za
Date: Tue, 20 Jun 1995 15:29:06 GMT+2
Subject: electropolishing platinum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In answer to Russel Cook's question on the above, I have found the
best solution is that of Tousek (Praktische metall. 7, 202, 1970)
which is:
equal volumes of nitric, sulphuric and phosphoric acids at about 3
volts ac.
I use a platinum strip as the other electrode. I normally pre-age the
solution for about two hours using two platinum electrodes at 3-3.5
volts ac. The solution turns from colourless to yellow. I keep the
solution temperature at 20 centigrade for pre-aging and polishing.
I have obtained conventional, HREM and ARM images using this method.


Dr MJ Witcomb
Electron Microscope Unit
University of the Witwatersrand
Private Bag 3
WITS
2050
South Africa

Telephone: + 27 11 716 4000
+ 27 11 716 2419 (messages)
Fax: + 27 11 339 3407
E-mail: mikew-at-gecko.biol.wits.ac.za




From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Tue, 20 Jun 1995 11:23:48 -0400 (EDT)
Subject: Job Opening/Technician

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

EM TECHNICIAN

A technician's position is available from July 3rd at The Electron Microscopy
Facility, College of Agriculture and Life Sciences, Cornell University for
teaching the laboratory portions of SEM and Freeze-fracture courses. The
applicant would be expected to spend a major part of the time training,
instructing, assisting and supervising students, postdocs and faculty on the
proper use of the equipment in the facility. Rest of the time would be
devoted to research service, maintaining equipment and up keep of the
laboratories. Applicants should have at least a B.S. degree in science,
preferably in biology. At least three years of experience in SEM, basic skills
in instrumentation and familiarity with computers is essential.
Familiarity with TEM techniques and Freeze-fracture is desirable but not a
requirement. The facility is well equipped and includes a Hitachi 4500 FESEM
and two Bal-Tec Freeze-fracture units. The applicant would work with
another senior technician in the Facility who would be the applicant's
immediate supervisor. Salary range is from $19,350 to $26,600.

Inquiries and applications should be directed to Dr. M.V. Parthasarathy at
the address indicated.

*************************************************************************
M.V. Parthasarathy
Professor of Plant Biology & Director, EM Facility
Section of Plant Biology
228 Plant Science Building
Cornell University, Ithaca, NY 14850
Telephone: (607) 255-1734
Fax: (607) 255-5407
E-mail: mvp2-at-cornell.edu
***********************************************************************




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Tue, 20 Jun 1995 13:02:00 -0400 (EDT)
Subject: Re: Legal/Ethical Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We routinely do work for local industry. Our University Legal Counsel
advised that as long as we did not use university resources to undercut
local competition we were O.K. It is also not a good idea to upset local
industry by undercutting their prices, because you may need their help
some day.
An example of the legal issues:

One of our microscope's service contract is payed for from NIH grant
support. It would be against our grant agreement and the law to charge
industry the same low rate we provide to academic users, because academic
users do not pay the full cost of instrument time (NIH grant picks up
those costs involved with service contract). Our rate to industrial users
MUST include compensation for part of the service contract. We also get
lights and electricity from the University. Our industrial rate must
include a component to compensate for these expenses.
Bottom line: you must make sure your rate reflects a level playing field
with other non-academic suppliers of similar services.

I hope this helps:

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************

On Mon, 19 Jun 1995, John C. Wheatley wrote:

} I am interested in the legal and ethical arguments (pro and con) for
} university involvement with industrial entities in the area of electron
} microscopy. In other words, is it legal/ethical for universities to use
} university instruments to solve problems for industry on a fee-for-service
} basis? I would like to have replies from anyone who wishes to respond.
} However, I realize that regulations may be different in other countries and
} therefore I ask that your responses be sent directly to me so our
} international friends are not bored with American legal/ethical
} discussions. I will provide a compilation of the responses to all who ask
} for them, except when individual respondents request otherwise.
}
} With respect to the legal side of the question, I would like to have
} answers based upon actual legal opinions that you are aware of.
} Perceptions about legality will not be useful. I know that there are
} probably very few (maybe none) lawyers involved in electron microscopy. If
} your institution has issued a legal opinion with respect to this issue, I
} would like to know what that opinion is.
}
} I would find it useful to have many people respond to this request. I
} would especially like to have those members of the private sector, who use
} electron microscopes to solve problems for industry, respond.
}
} If you feel comfortable responding, please include your price structure for
} industrial jobs--proprietary and otherwise.
}
} I thank you for your responses.
}
} John C. Wheatley
} Lab Manager
} Center for High Resolution Electron Microscopy--Arizona State University
}
} John C. Wheatley
} Lab Manager
} Center for High Resolution Electron Microscopy
} BOX 871704
} Tempe, AZ 85287-1704
} Phone: (602) 965-3831
} FAX: (602) 965-9004
} John.Wheatley-at-ASU.Edu
}
}
}




From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Tue, 20 Jun 1995 13:10:28 -0400 (EDT)
Subject: Re: Legal/ethical question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Bob Wise makes some good points about supporting local industrial
research not contributing to tenure decisions. However, one benefit that
can acrue from doing consulting EM work for outside industry is that the
monies generated often can be put into a discretionary fund you have
control over. This is a good way pay for lab equipment that Deans and
Government grants won't.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************





From: Michael Rock :      merock-at-u.washington.edu
Date: Tue, 20 Jun 1995 10:12:31 -0700 (PDT)
Subject: Re: Fe identification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Peter-
has anyone suggested using energy dispersive X-ray Spectrometery (EDS)
most analytical EM's are equipped with them. barring any Fe introduced by
sample prep or your grid (use Ni or Au grids) you should identify the Fe
in your sample. Depending on the granules and the resolution of your EDS
system you may be able to resolve whether the Fe is in the inclusions of
throughout the cytoplasm.
-Mike


On Mon, 19 Jun 1995, Dr. Molnar
Peter wrote:

} Hi Folks:
} Is there anyone out there who knows a specific method that is usable
} in conventional TEM for unequivocal identification of iron? I have a
} case of chronic granulomatous disease with Prussian blue deposition
} in macrophages. I would like to have a similar reaction done in EM.
} Any suggestions would be welcome.
} Thanks
} Peter Molnar M.D., Ph.D.
} Hungarian-Japanese EM Ctr
} Dept. Pathology, Univ. Med. Sch.
} Debrecen, POB. 24.
} Nagyerdei krt. 98.
} H-4012 Hungary, Europe
} FAX/Tel:36-52-417-063
} e-mail:molnarp-at-lib.dote.hu
}




From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Tue, 20 Jun 1995 13:38:21 -0400 (EDT)
Subject: Re: Legal/ethical question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


But if you have support (e.g. from your institution), beware. I know of a
library that had it's institutional $$ support reduced when contributions
(e.g. from alumni) made specifically for the library were increased. In
effect, the supposed additional income made no difference to the library's
ability to operate for that year.
-mc

} Bob Wise makes some good points about supporting local industrial
} research not contributing to tenure decisions. However, one benefit that
} can acrue from doing consulting EM work for outside industry is that the
} monies generated often can be put into a discretionary fund you have
} control over. This is a good way pay for lab equipment that Deans and
} Government grants won't.
}
} Jay Jerome







From: Michael Rock :      merock-at-u.washington.edu
Date: Tue, 20 Jun 1995 10:54:21 -0700 (PDT)
Subject: Re: Legal/Ethical Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John-
in response to legal/ethical.
I think they are separate issues and need to be tackled as such.
1) legal
I have been involved in such a system at a previous place of employment. We
operated as a "revenue generating" entity within the university, and set
up a system where we could do "service work" for outside clients. By doing
this we also had to charge our own users. We set inside rates to a
reasonable level so as not to discourage research, and the industrial
rates equal to or higher than other service labs in the area were
charging so as not to compete with them.
*we stayed legal*
2) ethical
this topic was much harder to define as to yes/no. there is a lot of gray
area out there. I was managing a electron microscopy cost center "revenue
generating facility" within the university, at a previous place of
employment. I agree with the philosophical argument, "to combine the
knowledge and resources of academia with the resources $ available in
industry, to solve practical problems and develop new technologies". BUT
there there are some very tough proprietary questions which need to be
dealt with in advance. Reasonable rates need to be established for faculty
and grad students so as not to discourage use, and higher rates need to
be established for the industrial clients, so there are no charges of
unfair competition from service labs in the area. Then there must be some
way of enforcing such policies on "tenured faculty" who may get lost in
this grey area "fog $$$".
Faculty sometimes had the idea that consulting jobs could be done at the
lower rate, or charged to their grants! This was very hard to monitor
and/or enforce, it relies on personal integrity of all those who are using
and administrating the facility.
good luck
-Mike




From: jandt-at-msc.cornell.edu
Date: Tue, 20 Jun 1995 15:52:43 -0400 (EDT)
Subject: to be forwarded (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Reply-To: jandt-at-msc.cornell.edu

Hi Folks,
}
}
} I would like to etch a PET (polyethylterephtalate) film in order to
} see the crystalline part protubating on the surface. My film is 180 microns
} thick, biaxially stretched. The crystallinity is approximatly 40%.
} I would prefer to etch it smoothly, to remove every layer one by one
} (i.e. without any crazing effect if possible). Does anyone know how to do that?
} Thanks a lot in advance.
}
}
jandt-at-msc.cornell.edu





From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Tue, 20 Jun 1995 15:28:16 -0500
Subject: AFM/DNA on carbon wafer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
A colleague of mine, Jim Sheldon, would like to adhere DNA onto a
carbon wafer for imaging with AFM. Anyone out there know of a relyable
protocol? Apparently, just adding the DNA to the wafer results in clumping
of the DNA. Thus, the carbon surface needs to be activated somehow.
Suggestions?
Thanks in advance,

Tobias Baskin

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
___ ____ ^ ____ _____ Tobias I. Baskin
/ \ / / \ / \ / University of Missouri
/ | / / \ / / Biological Sciences
/___ / /__ /_____\ / /__ 109 Tucker Hall
/ / / \ ( / Columbia, MO 65211 USA
/ / / \ \ / voice: 314-882-0173
/ /____ / \ \____/ /_____ fax: 314-882-0123






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Tue, 20 Jun 1995 13:57:37 -0700 (PDT)
Subject: LM:Used Zeiss WL wanted

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer01.u.washington.edu

Hello all,
We are looking to buy a used Zeiss WL. With external lamp, stage, and
objective turret. Eyepieces, objectives, camera tube and condenser are
optional. This will be used for brightfield microscopy and photomicrography.

TIA for any offers. Email me directly.


Glen MacDonald
Hearing Development Laboratories
University of Washington
Seattle, WA 98195-6515
(206)543-8360
glenmac-at-u.washington.edu







From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Tue, 20 Jun 1995 14:41:00 -0700 (PDT)
Subject: TEM:digital imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer01.u.washington.edu


Does anyone have some recommendations or warnings about digital image
capture from an TEM? I've been asked to help spec. out a system to be used
on a Philips EM 410.
I think the side mounted detectors with metal bellows have
chronic problems developing vacuum leaks.
What are experiences with 8-bit image depth versus 10-bit or
greater? Tentatively, I'm considering a Mac PowerPC 7100 with a Scion
AG-5 8-bit frame capture board, running NIH Image. Plus the usual
ethernet and optical drive.

Please email directly to cut the bandwidth on this group.

TIA
capture from TEM? I've been asked to help spec. out a system to go onto
a Philips EM410. Side mounted cameras seem to have chronic problems with
vacuum leaks developing in their metal bellows.
Any comments about 8-bit versus 10-bit or greater signal depth as
far as performance relative to cost? Tentatively, we are looking at a Mac
PowerPC (overkill, but fun) with an 8-bit framegrabber board and running
NIH-Image. And athe usual Ethernet and mass storage accessories.

TIA. Please send any commments directly to me.

Glen MacDonald
Hearing Development Laboratories
University of Washington
Seattle, WA 98195-6515
(206)543-8360
glenmac-at-u.washington.edu






From: GVKM07A-at-prodigy.com (DR CHARLES A GARBER)
Date: Tue, 20 Jun 1995 21:55:26 EDT
Subject: AFM/DNA on carbon wafer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

On June 20, Tobias Baskin mentioned "carbon wafers" as substrates for
AFM.

People seem to be using HOPG (highly ordered pyrolytic graphite) for
this purpose. It comes in several different grades, and there is
considerable variation in price between them.

If you want further information about the HOPG and the different grades
for this application, then we will be happy to send you our product
sheet on the product (which we sell, of course!) just for the asking.
Be sure to send a FAX number for fastest response. For fastest
response, use {SpiSupp-at-aol.com} .

The most expensive grade is for instrument calibration. To use this
"premier" grade for normal use, for most researchers, this would be
overkill. The grade STM-1 is quite acceptable for almost any high
quality research work and our grade STM-2, which we call "student"
quality is apparently quite acceptable for many people as well, but is
is much cheaper. The HOPG is somewhat like mica in that thin layers
can be "stripped" off. So one can get quite a few "strippings" from a
single block of the product.

Chuck

======================================================
Charles A. Garber, Ph. D.
President
SPI SUPPLIES
PO BOX 656
West Chester, PA 19381-0656 USA

Ph: 1-(610)-436-5400
1-(800)-2424-SPI

FAX: 1-(610)-436-5755

e-mail: GVKM07A-at-prodigy.com
SpiSupp-at-aol.com [SPI Customer Service e-mail box]
======================================================





From: Mnr HJ Els :      HJELS-at-op1.up.ac.za
Date: Wed, 21 Jun 1995 11:39:35 GMT+2
Subject: immuno gold labelling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Which is the best detergent to use in electronmicroscopical immunogold
laballing of intraerythrocytic parasites? ?time; ?concentration?




From: Mnr HJ Els :      HJELS-at-op1.up.ac.za
Date: Wed, 21 Jun 1995 12:28:48 GMT+2
Subject: immuno gold labelling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
we are trying to improve immuno labelling.
which is the best detergent to use in electron microscopical
immunogold labelling of intra-erythrocytic parasites (babesia)?
what concentrations can be used?
how long should the detergent be applied (after fixation)?
any suggestions would be greatly appreciated.

E-mail: hjels-at-op1.up.ac.za




From: Richard E. Edelmann :      EDELMARE-at-CASMAIL.MUOHIO.EDU
Date: Wed, 21 Jun 1995 08:04:29 -500
Subject: Re: AFM/DNA on carbon wafer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

microscopy-at-aaem.amc.anl.gov

Tobias,
As Chuck mentioned you will need to use HOPG as the carbon
substraight. I haven't tried DNA yet, but I would suggest using the
Kleinschmidt DNA spreading technique to minimize clumping (sorry I
can't seem to find a reference for the technique presently but it is
very common for TEM preps). Basically the DNA is prepared in a
volatile buffer of ammonium acetate, and mixed with a protein
(cytochrome c), and the mixture is gently applied to to a clean water
surface. The denatured cytochrome c forms a film at the air/water
interface to which whole lengths of DNA as attached. This film could
then be picked up with a freshly cleaved sheet of HOPG (as apposed to
TEM grids, or mica).

Another method I remember comming across which in my mind held
great promise was floating the DNA on a drop of Mercury. As the
mercury is reduced in volume the organic molecules accumulate at the
apex of the drop. Then observation can be made with STM (conductive
mercury). I origianlly saw this technique discribed about 2 years
ago.


Richard E. Edelmann
Electron Microscopy Facility Supervisor
352 Biological Science Building
Miami University, Oxford, OH 45056
Ph: 513-529-5712
E-mail: edelmare-at-muohio.edu




From: Diane Montpetit :      MontpetitD-at-EM.AGR.CA
Date: Wed, 21 Jun 1995 07:54:02 -0400
Subject: mocha image analysis many thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {sfe7da6d.009-at-EM.AGR.CA}
X-Mailer: Novell GroupWise 4.1

many thanks from those who answered my question, I really appreciate the support.

I will certainly try all of the suggestions.

This microscopy group is a blessing ...Thanks Nestor.





From: ubira-at-iqm.unicamp.br (Ubirajara Pereira Rodrigues Filho)
Date: Wed, 21 Jun 1995 09:40:51 -0300
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Can someone help to obtain the e-mail of SPI supplies or Dr Charles
Garber ?
Thanks in advance,

Ubirajara Pereira Rodrigues Filho

e-mail: ubira-at-iqm.unicamp.br





From: EMLAB-at-opus.mco.edu
Date: Wed, 21 Jun 1995 09:21:52 -0500 (EST)
Subject: Re: immuno gold labelling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Wed, 21 Jun 1995 08:42:05 -0500
Subject: WWW Icons for microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to announce a "competition" for microscopy related icons
that can be used on the Web. (I am afraid that the only prize will be the
adoption of particular icons by people who develop home pages.) I can
accept icons using anonymous ftp to risc1.numis.nwu.edu - please place
any icons in the directory contrib. I will also accept icons if they
are emailed directly to me as ASCII text. Please note: all icons must
be in standard ASCII or UNIX form, not encrypted Mac language.
As contributions come in I will place them in a directory which
can be accessed via the Web to our homepage at
http://risc1.numis.nwu.edu/internet/lab.html .




From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Wed, 21 Jun 1995 08:49:45 -0500
Subject: HREM Shareware

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Within our homepage (http://risc1.numis.nwu.edu/internet/lab.html) I
am going to start collecting shareware for HREM. The primary categories that
I am thinking of are:
a) Semper (5 or 6) library programs or run programs.
b) X-windows based routines such as simple CTF graphers.
c) Other X-windows type programs.

I can take PC based routines, with many reservations Mac-based programs
but since we are primarily unix/workstation focussed this will be the exception.

Contributions can be sent via anonymous ftp to the directory contrib
at risc1.numis.nwu.edu; any requests for further information or clarification
can be emailed to me.




From: jandt-at-msc.cornell.edu
Date: Wed, 21 Jun 1995 09:52:54 -0400 (EDT)
Subject: PET etching

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Reply-To: jandt-at-msc.cornell.edu


} Hi Folks,
} }
} }
} } I would like to etch a PET (polyethylterephtalate) film in order to
} } see the crystalline part protubating on the surface. My film is 180 microns
} } thick, biaxially stretched. The crystallinity is approximatly 40%.
} } I would prefer to etch it smoothly, to remove every layer one by one
} } (i.e. without any crazing effect if possible). Does anyone know how to do that?
} } Thanks a lot in advance.
} }
} }
} jandt-at-msc.cornell.edu
}





From: hukee.margaret-at-mayo.EDU (Marge Hukee)
Date: Wed, 21 Jun 1995 10:23:24 +0200
Subject: detergent for immunogold

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506211521.AA18901-at-fermat.Mayo.EDU}
X-Sender: hukem-at-fermat.mayo.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

In our facility, Tween-20 has been used extensively in both PBS and TBS in
concentrations of .1% for post-embeddeding labeling procedures . But it
sounds as if you are using a pre-embedding technique. Maybe more info is
needed
Marge

Marge Hukee
EM Core Facility hukee.margaret-at-mayo.edu
Mayo Foundation (507) 284-3148
----------------------------------------------------------------------------
--






From: X.m. Burany :      burany-at-sfu.ca
Date: Wed, 21 Jun 1995 10:57:14 -0700 (PDT)
Subject: copy of paper

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can anyone out there can send me copies of the following paper by Fax.

Schlesinger, M.; Meng, Xianying; Snyder, D.D., J.Electrochem. Soc. 1990,
137(6) 1858-9.

Schlesinger, M.; Meng, X. Evans, W.T.; Saunders,D.A. Kampert, W.P., J.
Electrochem. Soc. 1990, 137(6) 1706-9.

Schlesinger, M.; Meng, Xianying; Snyder, D.D., J. Electrochem. Soc. 1991,
138(2) 406-10.

My Fax No. is (604) 291-3592.

I greatly appreciate your help.

Sandy Burany, PhD


Dept. of Physics
Simon Fraser University
Burnaby, B.C. V5A 1S6
Canada
burany-at-sfu.ca




From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Wed, 21 Jun 1995 13:51:19 -0700
Subject: Re: Digital TEM Imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Glen MacDonald asked about Digital Imaging on a TEM. I hope he realizes that
} the system he is looking at will be useless for producing quality images, and
} will have as its only advantage speed. Even an expensive CCD system (
}
}
}
}
}
} 1024 x
} 1024 x 14 bit) cannot equal film for image quaility, and a TV capture board
} is far below that in quality. Glen needs to carefully define his expectations
} before he goes about specifying a system.
} John Mardinly
} Intel Materials Technology

I would second John's comments. I have been doing some initial
evaluations, and the 1K X 1K images just don't seem to have enough
information to be useful for publication. I'm still waiting to hear from
the folks who were going to show me some images from a 2K X 2K camera.
Perhaps we will see some new capabilities at the MSA meeting in Kansas City
this August.

John
chandler-at-lamar.ColoState.EDU
http://www.vetmed.colostate.edu/anatomy/faculty/chandler.html






From: sulovsky-at-sci.muni.cz (Petr Sulovsky)
Date: Thu, 21 Apr 1994 17:53:50 -0400
Subject: PE tubing for INTEGREX Colourjet Printer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506211556.AA09849-at-elanor.sci.muni.cz}
X-Sender: sulovsky-at-elanor.sci.muni.cz
X-Mailer: Windows Eudora Version 1.4.3
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi folks,

could anyone help (directly or by advice) with finding plastic (PE probably)
tubing for ink circulation in an old (1988) INTEGREX Colourjet 132 ink jet
printer? Mine has been fused by a hot IC and the company servicing LINK /
OXFORD / CamScan products here is unable to replace the molten tubing.

Thanks in advance,

Petr Sulovsky







~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

U M M U
U M M M M U Dr. Petr Sulovsky
U M M M M U Dept. of Mineralogy,
Petrology and Geochemistry
U M M M M U Faculty of Science
U M M M M U Masaryk University
U M M M M U Kotlarska 2
U M M M M U 611 37 BRNO
U M M M M U Czech Republic
U M M M M U Phone + 42-541129231
U M MM M U Fax + 42-541211214
U M M M U e-mail: sulovsky-at-sci.muni.cz
U U
UUUUUUUU

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: Larry Allard :      allardlfjr-at-ma160.ms.ornl.gov
Date: 21 Jun 1995 16:51:05 U
Subject: Digital TEM imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1408368857.60343-at-ma160.ms.ornl.gov}

Subject: Time: 4:06 PM
OFFICE MEMO Digital TEM imaging Date: 6/21/95

Regarding John Mardinly's recent post:
John and I have been arguing about motorcycles longer than I care to
remember, so now maybe we can now have a new topic for argument:

Digital cameras offer many advantages over film usage in the electron
microscope. For high magnification, high resolution imaging, they offer
equivalent pixel resolution to the best photographic films. They provide
much higher dynamic range in the image, and they have essentially a linear
response function. They allow immediate processing of the image data so that
on-line decisions can be made regarding the data being taken (e.g. FFTs
provide information on the periodicities recorded in the image). However, if
you routinely take low magnification micrographs that cover large image
areas, and you enlarge those micrographs 20 times or more to make a poster,
you will have to "scale" the image (a function provided by e.g.
DigitalMicrograph from Gatan) to interpolate between pixels to avoid a
pixelated image (that looks like a bad image, but really is not). Edgar
Voelkl in my laboratory can provide an even better algorithm for image
scaling, as a plug-in for DM, for those who are intereste.

We have not had a single piece of film in our Hitachi HF-2000 in more than 2
years, and have not missed the hassle of film usage one single bit in that
period. I would not be surprised if a large majority of the laboratories
presently using digital imaging on their electron microscopes have
essentially eliminated film usage completely for the convenience of digital
imaging. Remember, most publications routinely degrade your images anyway,
so your best film images end up looking pretty bad when they finally are
published.

Larry Allard







From: GVKM07A-at-prodigy.com (DR CHARLES A GARBER)
Date: Wed, 21 Jun 1995 13:05:51 EDT
Subject: Legal/Ethical Questions for listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

In response to the original posting by John Wheatley and the replies by
others now totalling seven:

Everyone answering thus far has been employed by some major research
university, in charge of major tax-payer funded instrumentation, in
most instances, granted by the federal government, probably on an NSF
grant of one type or another (or in some cases, NIH). I now enter the
foray, not as a university faculty or staff member but as an owner of
one of those "service" laboratories referenced on more than a few
occasions. Just for the record, Structure Probe in an independent
laboratory offering laboratory analytical services to clients primarily
in industry. I am rather proud, in fact of what I did in 1970: With
private capital I risked everything that I had in this world and
against the advice of just about everyone who knew me at that time,
purchased the very first JEOL JSM U3 imported into the USA. My intent
was to offer a laboratory analytical service based on this new SEM.
Not one drop of government money, federal, state or otherwise, was used
in the formation of my firm. This "risk" investment also created jobs,
virtually from the very beginning, several of them of the higher paying
technical professional type.

And from this "seed" SEM, the company has grown to include the SPI
Supplies division and a work force of about 25 persons, many of them
highly trained professionals in the EM field.

I am personally astonished with the way one can rationalize as being
acceptable what is fundamentally an improper if not outright illegal
activity. This is nothing new to me since I have had to deal with this
"problem" now for over twenty five years. I have been invited to
testify on five different occasions before various committees of the
United State Congress to explain what it is like to be a small business
person in the United States and have as your main competitors
organizations who get their equipment given to them for free on a grant,
pay no corporate net income taxes at the federal, state, or local
levels, are free from their state's sales and use tax act, can import
equipment duty free, just to name a few of the unfair advantages. So
far I have not heard one single person say one word about the fairness
of it all. I have heard only that if the university charges
"commercial rates" then all will be OK and perceived to be "fair".

Another thing that astonishes me is that there has not been one mention
about NSF Important Notice 91. Now I don't know whether that is
because no one knows about it, or whether like 55 miles an hour speed
limits, people tend to ignore it, hoping they don't get caught. But
surely that should not be the policy of those employed in professional
positions of responsibility in some of our nation's leading
universities. Most of the commercial activities presented in the other
postings would be prohibited by NSF Important Notice 91 (I am less
familiar with NIH policies).

I have prepared a discussion on this issue but it is too long for a
posting on this listserver. Anyone interested in receiving what I have
prepared should e-mail me their request.

The bottom line, however, is that there should never be any situation
where a university's instrumentation would be used in the competitive
marketplace irrespective of what price is being charged. NSF Important
Notice 91 is very clear on that point. Price is not mentioned once
because price is irrelevant to the discussion. It would be a rare
instance, if ever, that it would be proper, legally or ethically, to
use equipment owned by any nonprofit or not-for-profit (or other type
of public entity), in competition with for profit tax-paying
organizations. No "spin doctor" can ever make it "kosher", it is just
plain wrong.

Chuck

======================================================
Charles A. Garber, Ph. D.
President
SPI SUPPLIES Div. of Structure Probe, Inc.
PO BOX 656
West Chester, PA 19381-0656 USA

Ph: 1-(610)-436-5400
1-(800)-2424-SPI

FAX: 1-(610)-436-5755

e-mail: GVKM07A-at-prodigy.com
SpiSupp-at-aol.com [SPI Customer Service e-mail box]
======================================================





From: BARRY PYLE :      umbbp-at-msu.oscs.montana.edu
Date: Wed, 21 Jun 1995 17:08:16 MDT
Subject: Image Analysis Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


June 21, 1995

Does anyone have information on an image analysis software package available
from (or at least developed for) the National Institutes of Health? I have
heard that this package is extremely good and best of all it should be in the
public domain or available for minimal cost. I believe the program was written
for use with MacIntosh based systems.

Hope to hear from someone! Barry Pyle, Microbiology Department,
Montana State University - Bozeman.




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Thu, 22 Jun 1995 17:03:38 +1100
Subject: Dye sub printers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Some 6 months to a year ago there was quite a lot of discussion on this
listserver about dye sublimation printers. We are hoping to purchase either
a slow-scan CCD camera or Accuview-type camera for a new TEM and need some
way of getting hardcopies as close as possible to photographic quality. I
still have most of the messages on this topic that were sent over the last
year but wondered if anyone would like to send me brief comments of their
own experiences with these printers. I suspect that the technology is
improving rapidly and what was state-of-the-art a year ago is probably now
superceeded. Your personal opinions or experiences (good or bad) with ANY
dye sub. printer models would be gladly received. I will treat any replies
which are sent directly to me confidentially.

Thanks in advance.

Richard E


Richard Easingwood
South Campus Electron Microscope Unit
Otago Medical School
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

"The southernmost electron microscope unit in the world"






From: Clint E. Young :      clintey-at-unixg.ubc.ca
Date: Wed, 21 Jun 1995 23:00:45 -0700 (PDT)
Subject: Macintosh Imaging Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Barry,

The best (free) imaging software package for the Macintosh is--without a
doubt--NIH Image written by Wayne Rasband.

The following is taken from the introduction of the NIH Image manual:

"NIH Image is a public domain image processing and analysis program for
the Macintosh. Briefly, it can acquire, display, edit, enhance, analyze,
print and animate images. It reads and writes TIFF, PICT, PICS and
MacPaint files, providing compatibility with many other applications,
including programs for scanning, processing, editing, publishing and
analyzing images. It supports many standard image processing functions,
including contrast enhancement, density profiling, smoothing, sharpening,
edge detection, median filtering, and spatial convolution...

Image supports Data Translation and Scion frame grabber cards for
capturing images or movie sequences using a TV camera. Acquired gray
scale images can be shading corrected and frame averaged."

You can acquire this freeware program via anonymous FTP:
ftp://zippy.nimh.nih.gov/pub/image/

-----------
There are, however, many other imaging software packages. For starters,
you should take a look at the following web page:

http://www.msi.umn.edu/SciVis/Packages/packages.html

Hope this helps.

Clint Young, B.Sc.
Department of Psychiatry
Universtity of British Columbia




From: sorin.lazarescu-at-scf.fundp.ac.be (Sorin Lazarescu)
Date: Thu, 22 Jun 1995 11:43:07 +0200
Subject: NIH Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




--========================_23134430==_
Content-Type: text/plain; charset="us-ascii"

"NIH Image" is the MacIntosh based image analysis software you asked
It is a shareware indeed and it requires a floating-point chip, or the
"Software FPU". You can dowlod it from any info-mac archive . This is one
of the many other lcations:

ftp://src.doc.ic.ac.uk/packages/info-mac/_Graphic/Utility/nih-image-152.hqx.gz

There is also an NIH Image mailing list. It was set up by a group in the Soil
Science Department at the University of Minnesota. To subscribe, send a
message containing the line "subscribe nih-image {your name} " to
soils.umn.edu.

In case if your mail has the attachment facility, double-click, in the
attachment message, on each of the followings items: "nih-image157.sea";
"nih-image157_docs.sea"; "SoftwareFPU 3.01." .



--========================_23134430==_
Content-Type: application/mac-binhex40; name="nih-image157.sea"
Content-Disposition: attachment; filename="nih-image157.sea"

(This file must be converted with BinHex 4.0)



--========================_23134430==_
Content-Type: application/mac-binhex40; name="nih-image157_docs.sea"
Content-Disposition: attachment; filename="nih-image157_docs.sea"

(This file must be converted with BinHex 4.0)



--========================_23134430==_
Content-Type: application/mac-binhex40; name="SoftwareFPU_3.01.sea"
Content-Disposition: attachment; filename="SoftwareFPU_3.01.sea"

(This file must be converted with BinHex 4.0)



--========================_23134430==_
Content-Type: text/plain; charset="us-ascii"

**********************************************************
Sorin Dan LAZARESCU e-mail:
sorin.lazarescu-at-scf.fundp.ac.be
F.U.D.P. dept de Physique-LASMOS phone : +32 (0)81 72 47 13
61, rue de Bruxelles +32 (0)81 72 47 18
B-5000 Namur BELGIUM fax : +32 (0)81 72 47 07
**********************************************************



--========================_23134430==_--





From: Bram Koster :      koster-at-schubert.biochem.mpg.de
Date: Thu, 22 Jun 1995 13:10:44 -0600
Subject: Digital TEM imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I second Larry's comments regarding the advantages of CCD cameras. Our
Phlips CM200FEG is also equipped with a cooled 1024x1024 CCD camera. We also
are
very happy with digital data recording and have not put in film in more than 2
years. Our CM200FEG is used for a specific application (automated tomography)
for which the digital images of the CCD are essential.

However, the 1024x1024 CCD camera is not a convenient recording device for ALL
applications when we compare it to film, which can record images with more than
10000x10000 pixels. For example, high resolution images of large specimen areas
are crucial for imaging large structures, as well as for judging, documenting,
optimizing, etc., (routine) preparations of (negatively stained) biological
molecules. Of course, one could partly overcome the small image size by a
'montage' of a spot-scan series of 10x10 CCD images into one large digital
image of 10000x10000 pixels. When you do this, you will find the required data
collection time, as well as the required diskspace for --ONE-- such a large
image (200 MByte), rather inconvenient compared to just making an exposure on
film.

As mentioned above, we use on our CM200FEG exclusively a CCD camera, but the
other three microscopes in our group heavily depend on image recording on FILM
of large specimen areas with high resolution.

On Jun 21, 4:51pm, Larry Allard wrote:
}
} Digital cameras offer many advantages over film usage in the electron
} microscope. For high magnification, high resolution imaging, they offer
} equivalent pixel resolution to the best photographic films. They provide
} much higher dynamic range in the image, and they have essentially a linear
} response function. They allow immediate processing of the image data so that
} on-line decisions can be made regarding the data being taken (e.g. FFTs
} provide information on the periodicities recorded in the image). However, if
} you routinely take low magnification micrographs that cover large image
} areas, and you enlarge those micrographs 20 times or more to make a poster,
} you will have to "scale" the image (a function provided by e.g.
} DigitalMicrograph from Gatan) to interpolate between pixels to avoid a
} pixelated image (that looks like a bad image, but really is not). Edgar
} Voelkl in my laboratory can provide an even better algorithm for image
} scaling, as a plug-in for DM, for those who are intereste.
}
} We have not had a single piece of film in our Hitachi HF-2000 in more than 2
} years, and have not missed the hassle of film usage one single bit in that
} period. I would not be surprised if a large majority of the laboratories
} presently using digital imaging on their electron microscopes have
} essentially eliminated film usage completely for the convenience of digital
} imaging. Remember, most publications routinely degrade your images anyway,
} so your best film images end up looking pretty bad when they finally are
} published.
}
} Larry Allard
}
}
} -- End of excerpt from Larry Allard



--

Bram

------------------------------------------
Dr.ir. Abraham (Bram) J. Koster
Max Planck Institute for Biochemistry
Department Molecular Structure Biology
Am Klopferspitz 18A
D-82152 Martinsried (near Munich), Germany
tel: (49) 89 - 8578 2632
fax: (49) 89 - 8578 2641
email: koster-at-schubert.biochem.mpg.de
------------------------------------------




From: GVKM07A-at-prodigy.com (DR CHARLES A GARBER)
Date: Thu, 22 Jun 1995 08:08:39 EDT
Subject: Legal/Ethical Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

Thank you for requesting a copy of what I drafted up but which clearly
was too long to put out on the listserver. I offer this document not to
step on anyone's toes or to get people upset with me but to offer my
views, because what I see happening and going on is just plain wrong.

Let me start out by saying I believe that I am very qualified to
comment on this question. In 1970 I resigned my job at DuPont and
without a single client "lined up" in advance and armed only with my
belief that the SEM was going to be the "growth instrument of the
decade", I leveraged everything I had, purchased the very first JEOL
JSM U3 imported into the USA (it was serial #1) and started my own
business. I hired people at professional levels, paid good
professional level wages and also my full share of taxes, and
contributed, I think in a meaningful way to the economic development of
the region I live if not also the nation.

So it came as a real surprise to find out, in 1970, that there were
those persons, primarily at universities, who thought that they could
use their "own" university EM equipment as if such facilities were
their own to use for their own private so-called "consulting"
businesses. In just about any other circumstance, doing this would be
called white collar stealing, for which people do hard time, however
in "university-speak" this is called "consulting" and is somehow
presented to the students as "admirable behavior". One has to wonder
about the professor as role model and the effect on students!

Yes, on a regular basis, I too am hired as a consultant, and am paid a
daily fee, and the principal ability for me to do that is what is
between my ears. Not in "university-speak" but in plain English, this
is what most of the world calls "consulting". But when I leave my
client with samples under my arm to be characterized on some of our own
in-house instrumentation, then the principal ability to do it involves
access to the very expensive EM instrumentation and that is called by
everyone else, a "laboratory service". In any language other than
"university-speak", there is a clear difference between "consulting"
and "laboratory services".

Now I don't think that one has to be a rocket scientist to appreciate
that such use of the university's facilities is fundamentally wrong,
not just because of the potential for interference with legitimate
educational objectives but also because of the unfair competition it
presents for small firms like my own. After all, how could any
legitimate business compete long term with a competitor who did not pay
any taxes, did not have to pay for their equipment (it was granted via
NSF or some other funding agency), was exempt from import duties, and
as if this was not enough, they had a myriad of other benefits not
available to any for-profit tax-paying firm, either large or small.

Now those on the other side of the fence have argued that people with
my perspective are trying to "stop universities from doing research and
educating students" but nothing could be further from the truth. If
some commercial firm wants to give money to a university department for
work that is going to contribute to educational objectives, that is,
work that is basic and fundamental in nature, suitable for publication
in a student's thesis, and is intended for prompt publication in a
reputable scientific journal, then that is a win-win-win situation for
everyone. Surely I have no objection to that kind of activity and I
don't know of anyone else who would either. Indeed I would be the
first in line if one needed someone to speak in favor of the need for
universities to maintain a top notch ability to educate the researchers
of tomorrow, which can only be done if a top notch ability to conduct
research is maintained.

The objection is strictly to work being done that falls outside of that
which enhances educational objectives. The objection is with work that
either is not suitable for publication at all or else is proprietary in
nature and the client firm just would not want to see it published.
Why? Because he does not want to lose his competitive advantage and
give away his results to his competition. Unreasonable? I sure
wouldn't think so, after all he has a responsibility to the
shareholders to protect the company's assets and intellectual knowledge
is certainly as important an asset as any. So this then, by definition,
is the kind of work that more appropriately ought to be getting done in
the private for-profit sector and not in a university laboratory.

With the passing of time, some university administrations found they
had to do something to put a better face on what was going on in some
of their laboratories, and this led to some really creative approaches,
resulting in the formation of "industrial affiliate" programs, or
"industrial liaison" programs. Now I don't mean to paint all such
programs with the same brush, but I know of enough that are indeed
nothing more than shams to put a better face on what is basically an
inappropriate use of the equipment and instrumentation.

Let me give another example. A number of universities have set up not-
for-profit "research" institutes which becomes the official
"contractor" for outside users. The client companies contact with the
research institute, and the faculty for their consulting have their
consulting fees funnelled through the research institute. It then is
the research institute that, technically, does the contracting with the
university for access to the university's instrumentation. I am aware
of deans of engineering at such universities who will swear up and down
on a stack of bibles that "oh, no, we do absolutely no work directly
for private companies". They tell that to their state legislators.
They tell that to anyone who might ask. And some how, in their
convoluted way of thinking, they conclude that "oh, no, we are not
competing with private firms" and they think that such a sham kind of
structure makes them good citizens and operating within accepted norms.

In 1980, after complaints from numerous private testing and analytical
laboratories from across the USA, the NSF issued Important Notice 91
which essentially drew for the first time a line between what was and
was not appropriate usage of NSF funded instrumentation for outside
commercial users. A special committee, made up of top NSF officials
(the late Dr. George Pimentel played a key role in the formulation of
the final document as did also Dr. Donald Langenberg, now Chancellor of
the University of Maryland system), lawyers representing the NSF (Mr.
Charles Herz who is still at NSF in that same capacity), members of the
independent laboratory community (including myself) plus some other
knowledgeable people came up with the first draft of NSF Important
Notice 91. Some additional months passed before the document was ready
to be issued in its final form. And today, each and every NSF grantee
institution supposedly signs off each year certifying that they are in
full compliance with NSF Important Notice 91. I realize that NSF has
not done an especially good job in publicizing the existence of NSF
Important Notice 91, possibly the reason why none of the prior postings
on this subject have referenced Important Notice 91. But NSF Important
Notice 91, like it or not, has been the "law of the land" since its
release in 1980.

I decided to do a public rather than a private response to the John
Wheatley posting. Ethical values know no national boundaries, and in
any case, the issue is being discussed in other countries as well. None
of us should ever fear spreading the gospel of high ethical values and
conduct in the EM laboratory. Everyone does a lot of talking about
ethical values but did you ever try to write down a definition? The
dictionary defines "ethics" as "conforming to the standards of conduct
of a given profession or group". I like to extend that definition so
that when one talks of ethical values, it means that we are behaving
according to the definition EVEN WHEN NO ONE IS LOOKING AND THERE IS NO
CHANCE OF GETTING CAUGHT IF WE DON'T. In any case, behavior of faculty
in a university has to be held to a standard that is at the highest,
not the lowest common denominator. You have no idea what I have
learned from students over the years as to what goes on in some
departments. Or course students are caught between the rock and the
hard place: Sure they like the extra money that comes from running an
SEM or TEM for their advisor, when no one is looking, in the wee hours
of the morning. On the other hand, they have a good sense that
something is not quite right, they might not be able to exactly
understand or verbalize exactly why it is not right, but it is not
wasted on them that the activity being done is coming in the back door
when no one is looking. Is this really the message students should be
getting as they are being prepared for their future lives as a
professional? Like father, like son, some of the biggest academic
"entrepreneurs" in universities are themselves the students of other
"famous" entrepreneurs. Actually I am using the word "entrepreneur" as
used in the language of "university-speak". Real world entrepreneurs
take major financial risks. What kind of a financial risk does one
take when they accept samples and run them on their equipment as part
of their "consulting" business? What kind of risk does one take when
they "double dip", that is, are being paid by the university and for
the same hours worked, are being paid by the client company? Now maybe
this does not actually go on, maybe the "consulting" work is being done
on vacation time, or whatever, but it is the perception of the students
that count, isn't it? If perception is 98% of reality, the students,
or at least some of them, see their professors as something less than
honest people. If there are any doubts about this, stick around our
exhibit booth at MSA and listen for yourself what student have to say
when they drop off their resumes.

Another point about ethical values and that is that they are not the
kind of thing that you leave at the door when you walk into a different
room. Someone with high ethical values in one part of their life
probably has high ethical values in all parts of their life. The
conduct of high quality research demands honesty and that the
researchers have the highest of ethical values. One has to wonder to
what degree research results can be trusted when obtained by someone
who sneaks samples in the back door when no one is looking. Or when
such convoluted rationalizations are proposed so that somehow,
activities that are fundamental wrong and unfair, are somehow made to
seem right and fair.

I have not posted any prices and I certainly hope no one else does
either since that would come dangerously close to violations of the
anti-trust laws. I would respectfully suggest that no one even think
about posting "your price structure for industrial jobs - proprietary
or otherwise". We may not be lawyers, but surely we should know enough
to not be asking others to submit prices for the purpose of comparing
"competitive" prices. The threat of doing "hard time" for doing that
is usually enough to keep private companies from even thinking about
doing something like that! It should be enough to keep us from doing
that kind of thing as well.

The State of Wisconsin (Wise posting) can pass all the legislation it
wants and I realize Madison is some distance from the nation's capitol,
however it is not that far away that it can out run the intent of
federal law and regulation. Today, NSF Important Notice 91 is the law
of the land and it became necessary only after it became very very
clear that universities could not police themselves in terms of the way
some of their instrumentation facilities were being commercialized,
either officially or unofficially..

Another point has to do with the conclusions made by some that "we are
not in competition with private companies". It is really the fox
guarding the chicken coop saying the chickens are safe. No one seems
to consider that new would-be start up firms are literally preempted
from the market place because prospective entrepreneurs know full well
that an investment in such a firm would fail because they could not
compete. I myself have been turned down by leasing companies because
they ask "what if the University of XXXXXXX nearby decides to buy one
and compete with you". Of course I would fail and they know it. So I
can't get my financing. And the instrumentation does not get
purchased. And the jobs never get created. The university is always
going to be in competition when work outside of the scope of real
educational objectives is involved.

This is surely a subject that is of critical interest to anyone
involved with electron microscopy. Whether we are in academia,
industrial laboratories, or even vendors like myself, developing, doing
research, and manufacturing and distributing products to the industry,
we all have a vested interest in having a healthier rather than weaker
market place. Maybe some evening at MSA a meeting room could be found
where a panel discussion could be arranged, so that everyone might gain
a better appreciation for the other side's position on this subject,
with the ultimate goal being that we would be working together for our
mutual benefit on some of these issues.

Chuck

======================================================
Charles A. Garber, Ph. D.
President
SPI SUPPLIES / Division of Structure Probe, Inc.
PO BOX 656
West Chester, PA 19381-0656 USA

Ph: 1-(610)-436-5400
1-(800)-2424-SPI

FAX: 1-(610)-436-5755

e-mail: GVKM07A-at-prodigy.com
SpiSupp-at-aol.com [SPI Customer Service e-mail box]
======================================================





From: Jay Jerome :      jjerome-at-isnet.is.wfu.edu
Date: Thu, 22 Jun 1995 10:21:04 -0400 (EDT)
Subject: Re: Legal/Ethical Questions for listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Charles-
I would be very interested in receiving a copy of your posting. I will
supply it to our legal people here. For the time being, however, their
interpretation differs from yours.
As for the ethical issues. At many Medical Schools these days, we as
directors of laboratories are essentially independent entrepreneurs. My
laboratory only remains open as long as it generates enough capital to
pay the salaries and fringe of my technical people, service and operating
costs of my microscopes, and that portion of my time required to run the
facility. If this does not occur, I will be shut down. It is the same
with all of our support facilities (NMR, Flow cytometry, etc.). That is the
reality of the world these days. That I accomplish this using a mixture
of hospital service, grant support (many small businesses also use
government grants and contracts), and service to in-house and outside
is irrelevant as long as what I charged for that work represents fair
cost for doing the work. If I use University of Government money to pay
the service contracts and use the benefit to undercut prices THAT IS
UNFAIR, UNETHICAL, and ILLEGAL. If I generate business because I am close
to the user, have expertise he requires, or because he trusts my product
more than a competitors- that seems to me to be fair competition. As long
as the playing field is level I see no reason not to offer service to the
outside community. We don't do a lot of it, but what we do is important
to the financial well being of my facility. After all, many small
businesses regular try to get me to use them for my cell culture needs or
to produce monoclonal antibodies and oligonucleotides in direct
competition with our In-House tissue culture laboratory and probe
generating facilities.

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************





From: Greg :      GREG-at-umic.umic.sunysb.edu
Date: Thu, 22 Jun 1995 10:50:47 EST5DST
Subject: Digital Imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am interested in digitizing a JEOL 5300 SEM, Tungsten filament,
with a passive system. This is nonbeam control. The SEM is
in a core facility and both biological and materials are viewed. Has anyone
up gradedan SEM with after market equipment and software? Are you happy
with the results?

Greg-at-umic.umic.sunysb.edu
SUNY at Stony Brook
Univ. Microscopy Imaging Center





From: Sally Shrom :      sally-at-retina.anatomy.upenn.edu
Date: Thu, 22 Jun 1995 11:28:20 -0400 (EDT)
Subject: TEM/Bio/EPON

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English

Thanks to Stacie Kirsch at Electron Microscopy Sciences I have
finally solved my epon problem. I had trouble with my immuno experiments
when I used DAB with epon. I now know that the tissue looks great with
the epon from EMS. The problem is with the DAB ( not from EMS). Thanks
for all who responded to my original epon posting. I learned a lot...Sally




From: emccjb-at-ttuhsc.edu (Charles J. Butterick)
Date: Thu, 22 Jun 1995 11:28:20 -0400 (EDT)
Subject: Forensic/Balllistic Lab Needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
To: MICROSCOPY-at-AAEM.AMC.ANL.GOV

Greetings,
My apologies in advance to anyone who thinks my question is
inappropriate for this List.
Our local police asked me if I had the microscope that would enable
them to examine two bullets simeltaneously. I don't. They would send it to
Austin but Austin has a 6 month backlog and an answer is needed in 10 days
for a warrant.
Does anyone know of a lab does ballistics comparison with a
microscope? Please respond to me directly and not to the list.
Thanks.



Charles J. Butterick (Chuck)
Electron Microscopy Center
Department of Cell Biology
and Biochemistry
Texas Tech University Health
Sciences Center
3601 4th Street
Lubbock, Texas 79430

vox (806) 743-1633
fax (806) 743-1219
email emccjb-at-ttuhsc.edu or
chuck-at-micron1.lubb.ttuhsc.edu






From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-AAEM.AMC.ANL.GOV
Date: Thu, 22 Jun 1995 9:17:23 -0500 (CDT)
Subject: Do not post Software to the List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



G'day All

Yes, I saw the posting with the Software attachment
and have sent a note to the individual. Just as
a reminder to everyone. DONOT post code/programs/images
or any file longer than a few pages of text. If you
have something large that you wish to distribute
send it to me seperately and I will put it up
on the Anonymous FTP server which services the
Public Domain Software Libraries here at ANL.

You can then send your message with the information
that the data is on the FTP Server and anyone who
wants it can then retrieve it at their convenience.

Sorry, folks but I DONOT prescreen the postings.

Nestor
Your Friendly Neighborhood SysOp




From: ychen-at-macc.wisc.edu
Date: Thu, 22 Jun 1995 13:31:54 -0500
Subject: No big file on net, please!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {25062213272294-at-vms2.macc.wisc.edu}

Dear netter,
Please send big file off net. Otherwise it will jam the whole net.
Thank you!

Ya Chen



Ya Chen

=========================================================================
\ / Assistant Researcher-Cryo/SEM Coordinator TEL : 608-263-8481
\ / __ Integrated Microscopy Resource (IMR) FAX : 608-265-4076
\/ / / University of Wisconsin-Madison
/ / / 1675 Observatory Drive #167 Email:YChen-at-macc.wisc.edu
/ /__/_ Madison, WI 53706 Email:chen-at-calshp.cals.wisc.edu
=========================================================================






From: sorin.lazarescu-at-scf.fundp.ac.be (Sorin Lazarescu)
Date: Thu, 22 Jun 1995 18:40:17 +0200
Subject: Apologize

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am terribly sorry for the inconveniences I've caused to all of you, today.
That happend when sending, (to someone, in private), some image analysis
software, I forgot to disable the { {Cc: microscopy-at- aaem.amc.anl.gov} } .
I have no excuse and, once more, I am very confused and I apologize for the
mess I've done to all of you.

**********************************************************
Sorin Dan LAZARESCU e-mail:
sorin.lazarescu-at-scf.fundp.ac.be
F.U.D.P. dept de Physique-LASMOS phone : +32 (0)81 72 47 13
61, rue de Bruxelles +32 (0)81 72 47 18
B-5000 Namur BELGIUM fax : +32 (0)81 72 47 07
**********************************************************






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Thu, 22 Jun 1995 12:02:27 -0700 (PDT)
Subject: Re: Digital Imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer01.u.washington.edu

Greg,
We have been happy with results obtained by simply running a BNC
cable from our JEOL's video out port to an 8-bit framegrabber board in a
Macintosh computer. We capture with NIH-Image, timing the capture when
the scan reaches the bottom of the screen. This is for biological
samples, and has given publication quality images up to fairly high mags,
at least on our samples.
When we get enough other users interested, then we'll probably buy a system
that will give us full computer interface/control of the SEM.
I could send you the tiff files.

Regards,
Glen

On Thu, 22 Jun 1995, Greg wrote:

} I am interested in digitizing a JEOL 5300 SEM, Tungsten filament,
} with a passive system. This is nonbeam control. The SEM is
} in a core facility and both biological and materials are viewed. Has anyone
} up gradedan SEM with after market equipment and software? Are you happy
} with the results?
}
} Greg-at-umic.umic.sunysb.edu
} SUNY at Stony Brook
} Univ. Microscopy Imaging Center
}
}




From: Greg :      GREG-at-umic.umic.sunysb.edu
Date: Thu, 22 Jun 1995 15:17:16 EST5DST
Subject: Re: Digital Imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Date: Thu, 22 Jun 1995 12:02:27 -0700 (PDT)
} From: Glen Macdonald {glenmac-at-u.washington.edu}
} To: Greg {GREG-at-umic.umic.sunysb.edu}
} Cc: microscopy-at-aaem.amc.anl.gov
} Subject: Re: Digital {SEM} Imaging

} Greg,
} We have been happy with results obtained by simply running a BNC
} cable from our JEOL's video out port to an 8-bit framegrabber board in a
} Macintosh computer. We capture with NIH-Image, timing the capture when
} the scan reaches the bottom of the screen. This is for biological
} samples, and has given publication quality images up to fairly high mags,
} at least on our samples.
} When we get enough other users interested, then we'll probably buy a system
} that will give us full computer interface/control of the SEM.
} I could send you the tiff files.
}
} Regards,
} Glen
}
} ?
} }
} } Greg-at-umic.umic.sunysb.edu
} } SUNY at Stony Brook
} } Univ. Microscopy Imaging Center
} }
Glen,
Please send the TIFF images. It is good to see how these
systems work in the real world.--Thanks, Greg
} }
}





From: jfmjfm-at-umich.edu (John F. Mansfield)
Date: Thu, 22 Jun 1995 11:03:22 -0400
Subject: Re: HREM Shareware

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: jfmjfm-at-srvr5.engin.umich.edu
Message-Id: {v02120d0dac0f2cf8d75a-at-[141.213.21.13]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

It strikes me that this is entirely counterproductive to the effort that we
have put in over many years maintaining the Electron Microscopy and
Microanalysis Public Domian Library and the Microbeam Analysis Software
library. try the following:

ftp://www.amc.anl.gov/AMC-3/ANLSoftwareLibrary/
ftp://freebie.engin.umich.edu/pub/MSA+MAS/

Those wishng to contribute should send mail to zaluzec-at-aaem.amc.anl.gov for
the EMMPDL and to john.f.mansfield-at-umich.edu for the MASSL.


} Within our homepage (http://risc1.numis.nwu.edu/internet/lab.html) I
} am going to start collecting shareware for HREM. The primary categories that
} I am thinking of are:
} a) Semper (5 or 6) library programs or run programs.
} b) X-windows based routines such as simple CTF graphers.
} c) Other X-windows type programs.
}
} I can take PC based routines, with many reservations Mac-based programs
} but since we are primarily unix/workstation focussed this will be the
} exception.
}
} Contributions can be sent via anonymous ftp to the directory contrib
} at risc1.numis.nwu.edu; any requests for further information or clarification
} can be emailed to me.

John Mansfield
North Campus Electron Microbeam Analysis Laboratory
413 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX (313)936-3352
jfmjfm-at-engin.umich.edu, John.F.Mansfield-at-umich.edu
or jfmjfm-at-umich.edu they all reach me!
URL: http://www.engin.umich.edu/~jfmjfm/jfmjfm.html






From: bell-at-medcolpa.edu (Ted Bell)
Date: Thu, 22 Jun 1995 17:49:47 -0400
Subject: need to replace an outdated bioquant system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We currently have an old bioquant system capable of flourescence and want to replace it with a state of the art system, which is not from Bioquant. First off, are there any recommendations, especially in terms of cameras and frame grabbers?

1) Currently, the computer is 33MHz Gateway 486? Is it worth replacing with a Pentium or second-generation PowerMac?

2) Which kind of camera is best, digital or analog? What the are the advantages and disadvantages of each?

3) If we go with an analog camera and frame grabber, does anyone know if any company is now offering PCI frame grabber cards for the second-generation Power Macintosh yet?






From: gfan-at-ucsd.edu (by way of jpawley-at-macc.wisc.edu (James Pawley))
Date: Thu, 22 Jun 1995 12:21:16 -0600
Subject: Re: Digital TEM imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506221701.AA19386-at-sage.macc.wisc.edu}
X-Sender: jpawley-at-vms2.macc.wisc.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Nice balanced answer Gary. I would only add that this reduction in data
} resolution (below what one might expect from the "pixel area" of the CCD),
} is a strong function of voltage (higher is worse).
}
} One can use thinner phosphors to limit the effect but this reduces the
} signal level and is not too effective if fiber-optica couplings are used
} because the FO scatters many of the electrons passing down through the
} phosphor back up through it again, on a different trajectory.
}
} Likewise the dynamic range of the entire detector is far less that that of
} the (cooled?) CCD camera itself. The phosphor thickness is usually chosen
} so that each beam electron produces, say 10 electron/hole pairs in the CCD.
} This reduces the CCDs dynamic range by 10x.

I agree with you on these points.

} If you really want wide dymanic range, and digital imaging with a faster
} record-cycle time (between recorded images) than a CCD (which can take many
} seconds to read out), try a Fuji Image Plate system...
}
} Jim Pawley

and wait for a few days to see the images. Incidentally, there will be
a talk on the imaging plate at the symposium I mentioned.

Yes, readout time is a problem at the present. We are trying
a readout scheme that is much faster.

Just remind you that you replied to me, not the list.

Gary Fan
UC San Diego








From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Thu, 22 Jun 1995 17:55:41 -0500
Subject: Re: HREM Shareware

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The purpose of my email was not an attempt to kill either
EMMPDL or MASSL but completely different. The idea is to collect
primarily X-based routines for HREM and other miscellaneous such as
Semper code and libraries. I believe that there has been a disturbing
trend over the last ten years for heavy-duty programs for HREM to
be written on Government funding then sold commercialy. These programs
should be much easier and cheaper than they are (} $5000 in many cases).
I intend to donate NUMIS to this (at least the multislice and imaging
parts) as soon as I can rewrite for X-windows and thereby escape a
licensing agreement.

The more sites, the more software, the better!




From: Elinor Solit :      cambrex-at-world.std.com
Date: Thu, 22 Jun 1995 17:28:02 +0059 (EDT)
Subject: Re: Forensic/Balllistic Lab Needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chuck,

Any of the four light microscope manufacturers can probably do this for
you on an application basis. Numbers at hand are:

Carl Zeiss (ask for Ernest Keller) 800-356-1090
Olympus 800-446-5967
Leica 800-248-0123

Hope this helps.

Ellie Solit
"The Microscope Book" Call me if more assistance needed. 800-440-0311

On Thu, 22 Jun 1995, Charles J. Butterick wrote:

} Greetings,
} My apologies in advance to anyone who thinks my question is
} inappropriate for this List.
} Our local police asked me if I had the microscope that would enable
} them to examine two bullets simeltaneously. I don't. They would send it to
} Austin but Austin has a 6 month backlog and an answer is needed in 10 days
} for a warrant.
} Does anyone know of a lab does ballistics comparison with a
} microscope? Please respond to me directly and not to the list.
} Thanks.
}
}
}
} Charles J. Butterick (Chuck)
} Electron Microscopy Center
} Department of Cell Biology
} and Biochemistry
} Texas Tech University Health
} Sciences Center
} 3601 4th Street
} Lubbock, Texas 79430
}
} vox (806) 743-1633
} fax (806) 743-1219
} email emccjb-at-ttuhsc.edu or
} chuck-at-micron1.lubb.ttuhsc.edu
}
}




From: Michael Rock :      merock-at-u.washington.edu
Date: Thu, 22 Jun 1995 09:24:24 -0700 (PDT)
Subject: Re: Dye sub printers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Richard-
we have a Tektronix "Phaser II SDX" dye sub machine, the prices have come
way down over the past 2-3 years, the images are quite nice, and the
print speed is adequate. the only problems we have experienced are with
the mechanical paper feed, we get frequent paper jams. kind of annoying,
but with extensive cleaning, can be minimized.
-Mike

On Thu, 22 Jun 1995, Richard Easingwood wrote:

} Some 6 months to a year ago there was quite a lot of discussion on this
} listserver about dye sublimation printers. We are hoping to purchase either
} a slow-scan CCD camera or Accuview-type camera for a new TEM and need some
} way of getting hardcopies as close as possible to photographic quality. I
} still have most of the messages on this topic that were sent over the last
} year but wondered if anyone would like to send me brief comments of their
} own experiences with these printers. I suspect that the technology is
} improving rapidly and what was state-of-the-art a year ago is probably now
} superceeded. Your personal opinions or experiences (good or bad) with ANY
} dye sub. printer models would be gladly received. I will treat any replies
} which are sent directly to me confidentially.
}
} Thanks in advance.
}
} Richard E
}
}
} Richard Easingwood
} South Campus Electron Microscope Unit
} Otago Medical School
} PO Box 913
} Dunedin
} NEW ZEALAND
}
} Telephone: 64-03-479 7301
} Facsimile: 64-03-479 7254
}
} "The southernmost electron microscope unit in the world"
}
}
}




From: jpawley-at-macc.wisc.edu (James Pawley)
Date: Thu, 22 Jun 1995 18:38:19 -0600
Subject: Re: Digital TEM imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506222318.AA20720-at-sage.macc.wisc.edu}
X-Sender: jpawley-at-vms2.macc.wisc.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Nice balanced answer Gary. I would only add that this reduction in data
} resolution (below what one might expect from the "pixel area" of the CCD),
} is a strong function of voltage (higher is worse).
}
} One can use thinner phosphors to limit the effect but this reduces the
} signal level and is not too effective if fiber-optica couplings are used
} because the FO scatters many of the electrons passing down through the
} phosphor back up through it again, on a different trajectory.
}
} Likewise the dynamic range of the entire detector is far less that that of
} the (cooled?) CCD camera itself. The phosphor thickness is usually chosen
} so that each beam electron produces, say 10 electron/hole pairs in the CCD.
} This reduces the CCDs dynamic range by 10x.

I agree with you on these points.

} If you really want wide dymanic range, and digital imaging with a faster
} record-cycle time (between recorded images) than a CCD (which can take many
} seconds to read out), try a Fuji Image Plate system...
}
} Jim Pawley

and wait for a few days to see the images. Incidentally, there will be
a talk on the imaging plate at the symposium I mentioned.

Yes, readout time is a problem at the present. We are trying
a readout scheme that is much faster.

Just remind you that you replied to me, not the list.

Gary Fan
UC San Diego








From: Chris Jefferies :      chris-at-stowey.demon.co.uk
Date: Fri, 23 Jun 1995 00:03:14 GMT
Subject: Re: HREM Shareware

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In his message dated Thursday 22, June 1995 John Mansfield wrote :

} It strikes me that this is entirely counterproductive to the effort that we
} have put in over many years maintaining the Electron Microscopy and
} Microanalysis Public Domian Library and the Microbeam Analysis Software
} library.

I'm puzzled here. I don't see how an additional ftp site can be
*counter*productive.

Surely most of us would agree that an extra site spreads the load a little
and may well specialise in a slightly different area from existing sites.
This isn't to detract from the hard work (much appreciated BTW) that John M
and Nestor Z put into creating and maintaining their resources.

I already have Nestor's ftp site on the 'Microscopes and Microscopy' Web
pages. I'd missed John's (but I'll add it now). I'll certainly add the new
HREM site as well. The more the merrier I think. :-)

Regards to everyone from Bristol, UK

Chris
--
---------------------------------------------------------------------------
| Chris Jefferies E-Mail at home - chris-at-stowey.demon.co.uk |
| at work - chris.jefferies-at-bbsrc.ac.uk |
| Microscopy page {URL: http://metro.turnpike.net/jefferie/index.html} |
---------------------------------------------------------------------------




From: Roar Kilaas :      Roar_Kilaas-at-macmail7.lbl.gov
Date: 22 Jun 1995 18:12:29 U
Subject: Re: HREM Shareware

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1408277709.50174-at-macmail7.lbl.gov}
microscopy-at-aaem.amc.anl.gov
X-Mailer: Mail*Link SMTP/QM 3.0.0

Reply to: RE} } HREM Shareware

With respect to the comment about making an X-window image simulation package available, I would like to inform that such a program
already exists. It is called NCEMSS and is available from the National Center for Electron Microsocpy at the Lawrence Berkeley
National Laboratory. The program is written by Dr. Roar Kilaas and is currently available for the following platforms: SUN, SGI,
Dec Alpha (OSF & OpenVMS), DecStations, and the IBM RISC Series computers. All versions except the DecStation and IBM versions are
posted at the site:
http://ncem/NCEM/computer/software.html
The two missing versions will be added shortly.
Roar Kilaas
Phone : 510-486-4618
e-mail : Roar_Kilaas-at-lbl.gov





From: Alex Titkov :      ALEX-at-bunyip.ph.rmit.edu.au
Date: Fri, 23 Jun 1995 11:00:38 EST-10
Subject: SEM of fluoridated plastics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,

We have a customer who wants to see the internal surfaces of light
pipes made of various materials. The relevant features are about 100
nm. The materials are fluoridated plastics (like teflon), but some of
them are unstable even at 130C.

We sputter coated the samples with gold (various thicknesses between
30 and 90 nm) and viewed them in a conventional SEM (W filament) at
15-25 kV keeping the beam current as low as possible (down to 30 pA).

The problem is that the samples with thin coatings are unstable under
the electron beam whereas thick coatings seem not to form a continuos
layer of gold on the plastic, resulting in an orange peel appearance
(200-300 nm dia. gold "islands"?).

We would appreciate any ideas about imaging such samples.

Thank you.
---
Alex Titkov
Electron Microscopy Unit
RMIT Applied Physics
GPO Box 2476V
Melbourne VIC 3001
AUSTRALIA
alex-at-bunyip.ph.rmit.edu.au
Voice:(03) 660 2205
Fax: (03) 660 3837




From: MicroToday-at-aol.com
Date: Thu, 22 Jun 1995 21:00:56 -0400
Subject: Digital vs. Film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Groupus-
For those of you that receive our newsletter "Mircoscopy Today", I trust that
you will agree that the resolution, etc. of our micrographs and images are AT
LEAST as "good" as produced by ANY publication. And I continue to be very
interested in the topic of film vs. digital input..
As a side topic - should you be interested in producing printed copies of
your micrographs, etc., allow me to stress the importance of the printing
process:
1) An Al plate, with the image burned in from a negative, carries the ink to
the paper. A (large) quality press with a good operater is critical. If you
go to Econoprint, you will get good garbage. And high quality print shops
are expensive!
2) The image, from film or digital, is shot thru a dot-screen to produce
the negative. A 150, 180 or 200 line screen is appropriate for good work.
The higher, the better - assuming that the printing press can handle it. A
good press with a poor screen, or a good screen on a poor press, will result
in MUD.
As the result of a previous "thread" on this topic, I went to the
professional seperation house that I work with (NOT my printer) and asked
their advice - which was that 4" x 5" transparencies were the superior,
followed by 35 mm slides. Following were paper positives and digital. In
fairness to all, it is tough to "look" at the origional of a digital image
and compare it to the printed result.
With my continued interest in this topic, I presented the question to David
Scharf. Should you not be familiar with his work, I can only submit that he,
as a microcopist/photographer, has produced more nifty EM micrographs than
all of us together.
Dave advises that at the point a digital image is 48 bits deep, and in a 40
mbyte file, it generally becomes superior to film. Below that, film is
superior.
I fully appreciate that there are "computer jocks" out there that are very
excited over their results - and have opinions that will remain unchanged.
I, however, would like to continue to become "educated" on this subject and
would appreciate any contribution to my education.
Peace and All Kinds of Good Will,
Don Grimes, Microscopy Today







From: Michael Rock :      merock-at-u.washington.edu
Date: Thu, 22 Jun 1995 10:23:15 -0700 (PDT)
Subject: Re: Legal/Ethical Questions for listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to Charles Graber and Jay Jerome- You both have very valid
points, and good reasons for speaking out.

First Jay- saying that your legal people have a different interpretation
than Chuck's point of view, is about as significant as saying O.J.'s
lawyers have a different interpretation than the prosecution lawyers.
Let's address what is "right" or "wrong", and understand that change is
the only constant.

Let's deal with this Important Notice 91, does it have any teeth?
How do we enforce it? or are rules just made to be broken? Or is this law
out of date, do we need to revise it for the times?

And second Chuck- Jay is correct when he states that this type of govt.
supported is happening all the time, it is a way of doing business,
keeping university/hospital labs open. I have worked in and managed several
labs where during "lean times", if not for "outside" revenue generating
projects, the lab would have been shut down. This does not make it OK,
but utilizing resources is very important.

Which brings me to my pet peeve, resource management in the EM industry.
Universities need to keep up with the ever changing technology, but at what
price? In academia many faculty seem to "need" a new EM every couple of years,
with each technological breakthrough, and equip it with all the latest
analytical equipment, but seem to forget who is paying for this. What is
the universities mission? to teach?, or to satisfy the whims of some very
spoiled individuals by supplying them with a new toy every time they cry
out?
Out there in the real world (industry), companies purchase equipment
when they can afford it, not "justify it" (academic lingo). Most of the
EM service labs are probably using equipment that is a decade or two old,
not a year or two. Most can't afford service contracts, and most...go broke.
Sure companies in the IC industry can purchase the top of the line
equipment, but they can afford it, it's part of what they need to do to
stay in business. Maybe they should be the leaders in this academic/industrial
partnership by opening up their facility for educational purposes instead of
placing the burden on the tax payers to buy the equipment and performing this
"consulting service"???

-Mike




From: mager-at-unixg.ubc.ca (Mary Mager)
Date: Thu, 22 Jun 1995 21:29:36 -0800
Subject: Re: Digital SEM imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: mager-at-pop.unixg.ubc.ca (Unverified)
Message-Id: {v01510101ac0fff085617-at-[137.82.220.144]}
Mime-Version: 1.0
Content-Type: text/plain; charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable



} I am interested in digitizing a JEOL 5300 SEM, Tungsten filament,
} with a passive system. This is nonbeam control. The SEM is
} in a core facility and both biological and materials are viewed. Has anyone
} up gradedan SEM with after market equipment and software? Are you happy
} with the results?

} Greg-at-umic.umic.sunysb.edu

Greg,
The only commercial passive SEM digital system I know of is the Quartz PCI
system. It is a Windows-PC system for acquiring, processing and archiving
images from the SEM. It normally stores images in TIFF format but can
export in about 20 formats. I have had one for about two years on a Hitachi
S-2300 (simple) SEM and I must say the darn thing is addictive. No-one who
has used it once will use the SEM without using it. My film use has dropped
to one-tenth, since most people in my multi-user university lab just use
the laserprinter prints. The slow-scan image (up to 2500 X 2000 resolution)
on the super-VGA screen is very striking. You can annotate and do line and
angle measurments on the image, then store, print or photo-replay. The
latest version has a full image database to keep track of all the images
and find them by searches on many different topics or keywords.
=7F
My lab was used for initial testing of this system on a real SEM and so I
have followed the progress of this product since its inception. The
installers say they can interface to any SEM and have also hooked up to my
S-570 (vertical picture) and H-800 STEM. I have been happy with the product
and the students all love the full Windows compatability. You can cut the
images to the clipboard then drop them into Word or any other program. Now
I sell discs to carry away the TIFF ore JPEG files instead of film. Also,
since my S-2300 does not have any annotation, I can annotate the Quartz PCI
image then re-spool it through the SEM's photo CRT. In fact, it turns your
SEM into a Polaroid printer for any image that the PCI system can import
(see 20 formats above).=7F
NSC Hitachi carries this in Canada and I think NSA Hitachi does in the US.
Hope this helps,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Eng, UBC
309 - 6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648, fax: 604-822-3619
email: mager-at-unixg.ubc.ca






From: Wu Yujing :      ih36-at-jove.acs.unt.edu
Date: Fri, 23 Jun 1995 00:27:37 -0500 (CDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: norenbur-at-onyx.si.edu (Jon L. Norenburg)
Date: Thu, 22 Jun 1995 11:36:49 -0500
Subject: Re: Legal/Ethical Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506221536.LAA10282-at-onyx.si.edu}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear folks,
I am surprised that there is not already an avalanche of responses to Dr.
Garber. In large measure I agree with him as concerns ethics and yes,
there are a lot of hypocritical facades out there. Whether or not his
brush is a little or a lot too broad I don't know, I have no direct
experience with "commercial" use of university labs. I do know of
colleagues who would not be running service labs, and whose labs would be
shut down without the modest commercial income that goes exclusively to
operating funds for the lab. Without that income they would not be
training any students and many of my colleagues who do non-sexy (i.e., no
or minimal public funding) morphological research in this day and age would
not be using electron microscopy as a tool.

I am very sensitive to Dr. Garber's concern about "fairness." But, as is
so common nowadays, some people like to see things as black and white - so
Dr. Gardner implies that there cannot be any possible justification for a
not-for-profit (which, by the way, also employ people who spend money at
stores, pay taxes, buy houses and raise kids that eat apple pie) accepting
commercial contracts. Even your elected congressional representative can
scam all the for-profit book and speaking engagements that their time and
importance in office garners for them. However, at the present rate of
evolution in Washington DC, the absolutist reasoning used by Dr. Garber may
soon dictate that all government functions be privatized, because they all
compete with someone or the potential of someone to provide a service for
profit (perhaps we can even replace the military with private militias). I
hope that he is planning on privately educating all those students who no
longer have labs and can no longer use government or commercial money to
buy his services or those of SPI.

Ethics enters into this argument only as concerns legalities and norms of
society; it has no position that says the private interests of an
entrepeneur override those of other individuals. If society insists on
certain laws (or if industry has bought them from equally ethical public
servants) - too bad, we have to live with them. If people find ways to
evade restrictions legally - too bad, we'll just have to live with
university research institutes. Yes, if a university enters into a
contract with NSF it should honor the terms of that contract. Once that
contract is no longer binding, I see no ethical problem, in principle, with
a university lab making ends meet through commercial contracts - I do see a
problem with the facilities being used inappropriately for personal gain IF
that is in violation of university regulations or civil law.

Entrepeneurs are not the only gamblers; loyalty to a public facility can be
just as much of a gamble - moreover, ethics has nothing to say about which
type of gamble is morally superior!

This note has been composed on personal time on a personally owned computer
with public domain software...
--Jon


Jon L. Norenburg {norenbur-at-onyx.si.edu}
Invertebrate Zoology - MRC 163
National Museum of Natural History
Smithsonian Institution, Washington, DC 20560
Voice 301-238-3508, Fax 301-238-3361






From: f705geri-at-mbox.tu-graz.ac.at
Date: Fri, 23 Jun 1995 08:53:12 +0200
Subject: Dye sub printers/Imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We also have a Tektronix "Phaser II SDX" printing system connected to a
Mac 840AV with a Mac 17inch monitor. We are using Digital Micrograph for
image processing. The problem we have is that in many cases the colours
in the output on the printer differs significantly from the image on the screen.
Does any one known how to adjust/calibrate such a system ?

Geri


DI Gerald Kothleitner
Forschungsinst. f. Elektronenmikroskopie
Steyrergasse 17
8045 Graz, Austia





From: f705geri-at-mbox.tu-graz.ac.at
Date: Fri, 23 Jun 1995 09:51:59 +0200
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We also have a Tektronix "Phaser II SDX" printing system connected to a
Mac 840AV with a Mac 17inch monitor. We are using Digital Micrograph for
image processing. The problem we have is that in many cases the colours
in the output on the printer differs significantly from the image on the screen.
Does any one known how to adjust/calibrate such a system ?

Geri


DI Gerald Kothleitner
Forschungsinst. f. Elektronenmikroskopie
Steyrergasse 17
8045 Graz, Austia





From: Baoming Huang :      bmhuang-at-sunchem.chem.uga.edu
Date: Fri, 23 Jun 1995 10:19:38 -0400 (EDT)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe



* * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * *
Baoming M. Huang bmhuang-at-sunchem.chem.uga.edu
Department of Chemistry bmhuang-at-uga.cc.uga.edu
University of Georgia Tel: (706) 542-9453
Athens, GA 30602-2556 FAX: (706) 542-9454







From: Goldmarker-at-aol.com
Date: Fri, 23 Jun 1995 10:25:23 -0400
Subject: microbiology listserver?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of the existence of a similar microbiology listserver
similar to Nestor's well-managed system?

This is a great group, but I need to talk to some environmental or
fermentation microbiologists about methods to quickly measure biomass
viability and to quickly id microbes.

My internet access is via America Online at the moment.

Any help would be appreciated.

Regards,

Donald P. Cox, Ph.D.
GOLDMARK BIOLOGICALS
437 Lock St
PHillipsburg, NJ 08865
(908) 859-2631
(98) 859-2875-FAX
E-mail: goldmarker-at-aol.com




From: bell-at-medcolpa.edu (Ted Bell)
Date: Fri, 23 Jun 1995 11:59:01 -0400
Subject: need to replace an outdated bioquant system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We currently have an old bioquant system capable of flourescence and want to replace it with a state of the art system, which is not from Bioquant. First off, are there any recommendations, especially in terms of cameras and frame grabbers?

1) Currently, the computer is 33MHz Gateway 486? Is it worth replacing with a Pentium or second-generation PowerMac?

2) Which kind of camera is best, digital or analog? What the are the advantages and disadvantages of each?

3) If we go with an analog camera and frame grabber, does anyone know if any company is now offering PCI frame grabber cards for the second-generation Power Macintosh yet?






From: jeanne_barker-at-merck.com (Jeanne Barker)
Date: 23 Jun 1995 12:41:48 U
Subject: An atlas of the brain??

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I Have a Question regarding... An atlas of the brain??
6/23/95 11:46 AM
Hi,
I am beginning a new project involving morphology of the brain of the mouse,
particularly the cerebrum and the hippocampus. We will be doing transmission
E.M. My question is this--Can anybody recommend an atlas that shows T.E.M.
photographs of normal brain and perhaps some pathology?
Thank-you,
Jeanne






From: L. D. Marks
Date: 6/23/95 6:32 AM
Subject: Re: None

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1408220316.97921-at-macmail7.lbl.gov}

Reply to: RE} None

Laurie,

you could provide a link to
http://ncem.lbl.gov/NCEM/computer/software.html
for users to download NCEM's free software.

I think it makes more sense for us to keep the latest versions up on our server, rather than send you new versions as we make
changes (sometimes as often as daily).

Mike O'Keefe
Acting Head, NCEM
--------------------------------------

I am trying to collect programs for some of the more heavy-duty
aspects of electron microscopy, e.g. Multislice, Bloch-wave,
Image Processing (Maximum Entropy etc), Multibeam calculations,
Adaptive and Multifactor FFT's etc. I intend to place these on
our home page here at Northwestern, or I can include links to
other places where they can be found via anonymous ftp. All
this software will be Public Domain material, and I am primarily
thinking about Unix-based software with an X-windows based approach
to graphics so that it will run on different platforms.

Can you please pass this message on to one of your students
who can either send me directly via anonymous ftp any code that
you can donate or give me the appropriate link. You can email
me at ldm-at-apollo.numis.nwu.edu, but please use risc1.numis.nwu.edu
(129.105.122.66) for anonymous ftp (a different computer with
more space). The Web page link is
http://risc1.numis.nwu.edu/internet/lab.html
ftp://risc1.numis.nwu.edu

Thanks

Laurie Marks

P.S. Please pass this on to anyone else who you think might have
heavy-duty software that they are willing to contribute.






From: Dr Ross Mackenzie :      ross.mackenzie-at-Materials.oxford.ac.uk
Date: 23 Jun 1995 14:55:29 +0100
Subject: Lectureship in Materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"microscopy-at-aaem.amc.anl.gov" {microscopy-at-aaem.amc.anl.gov}
X-Mailer: Mail*Link PT/Internet 1.0.1

The following advertisement will appear in the scientific & educational press
in the near future.
Please draw it to the attention of appropriate colleagues.

Further information about this post and about the Oxford Materials Department
can be obtained from our WWW server. (http://www.materials.oxford.ac.uk/)

Ross Mackenzie

-----------------------------------------------------------------------

University of Oxford
Department of Materials in association with Trinity College or St Hilda's
College
University Lecturership in Materials

Applications are invited for the above post, tenable from 1 January 1996.
Stipend
according to age on the scale of 15154-28215 per annum. The successful
candidate
may be offered a tutorial fellowship, for which additional emoluments would
be
available, by either Trinity College or St Hilda's College under
arrangements described
in the further particulars.

Candidates should be able to teach widely within a broad four-year Materials
Science
syllabus and to carry out a program of research into the mechanical
properties of
materials.

Further particulars (containing details of the duties and full range of
emoluments
and allowances attaching to both the university and college posts) may be
obtained from
The Head of Department, Department of Materials, Parks Road, Oxford OX1 3PH,
UK
(FAX 01865 273738, Tel: 01865 273737). The closing date for applications is
31 July
1995.

The University exists to promote excellence in education and research, and
is an equal
opportunities employer.

--------------------------------------------
Dr Ross Mackenzie
University of Oxford
Department of Materials
Parks Road, Oxford OX1 3PH

Phone 01865 273708 or 01865 273693
FAX 01865 273789
--------------------------------------------




From: hmeekes-at-biosci.mbp.missouri.edu (Herman Meekes)
Date: Fri, 23 Jun 1995 09:17:15 -0500
Subject: Re: SEM of fluoridated plastics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On June 23 } Alex Titkov wrote:

} Dear Microscopists,
}
} We have a customer who wants to see the internal surfaces of light
} pipes made of various materials. The relevant features are about 100
} nm. The materials are fluoridated plastics (like teflon), but some of
} them are unstable even at 130C.
}
} We sputter coated the samples with gold (various thicknesses between
} 30 and 90 nm) and viewed them in a conventional SEM (W filament) at
} 15-25 kV keeping the beam current as low as possible (down to 30 pA).
}
} The problem is that the samples with thin coatings are unstable under
} the electron beam whereas thick coatings seem not to form a continuos
} layer of gold on the plastic, resulting in an orange peel appearance
} (200-300 nm dia. gold "islands"?).
}
} We would appreciate any ideas about imaging such samples.



I am not familiar with materials science, but I wonder if you could use
Pt/C replica techniques to solve the problem.

Briefly: expose the surface to be studied to a source of evaporating Pt in
high vacuum ( at a low angle). Stabilize by evaporating Carbon at a
perpendicular angle. Excise the parts of interest from your sample and free
the replica from the mould. View in TEM.
I seem to remember there even is a way of making an indirect replica by
first replicating the surface with a substance (resin?) and preparing the
Pt/C replica on that. Resolution in both instances should be in the 1-10 nm
range.

I think that more elaborate explanations of these techniques can be found
in EM manuals.




Herman Meekes
Biological Sciences ______________ ______________
University of Missouri ---__ \ / __---
109 Tucker Hall ------__\---/__------
Columbia, MO. 65211, USA \( )/
Tel: 314-882-0171 V
Fax: 314-882-0123 / \
e-mail: hmeekes-at-biosci.mbp.missouri.edu /___\






From: jpawley-at-macc.wisc.edu (James Pawley)
Date: Fri, 23 Jun 1995 12:14:17 -0600
Subject: Re: Legal/Ethical Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506231654.AA22089-at-sage.macc.wisc.edu}
X-Sender: jpawley-at-vms2.macc.wisc.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hear! Hear! to Jon Norenburg for taking the time to respond to some of the
points of Chuck Garber's diatribe. The thought of doing so wearied me that
I wanted to ignore it. Jon's example has woken me up!

He could have added that, at many respected institutions of higher learning
(Harvard, MIT ...), only about 50% of the prof's salaries are paid by the
institution and the prof is EXPECTED to garner the remaining 50% by
consulting. Are tuition rates not sufficiently high already? Pay them
less you say? How else does one get profs in technical fields to stay in
the university when they have to compete with manufacturers producing
military toys on "cost-plus" contracts? (Contracts, moreover, that are
increasingly used, in part, to pay EM service labs!)

Has Chuck heard about the almost total demise of the industrial microscopy
labs in the US? of the many 400kV, 300kV TEMs and VG-STEMs etc. given
away to anyone with a truck? Tax policies dictate this sort of shift but
someone still has to do the work, (unless we want it all done in Asia or
Europe). Why not the universities who have students to train and questions
to answer? The fact is that this system is often the most "cost effective"
way for a company to get an answer quickly and right. Like many economic
debates, it all depends on whatyou count as costs and what you are willing
to ignore to make the argument sound better.

We have heard a lot about NSF Doc. 91 but the microscopes in universities
and other non-profit organizations are not just bought using NSF funds.

We have heard about the Profs who just must replace their EMs every year or
two. I don't know where these profs are. Comprehensive surveys by MSA and
others show that the stock of microscopical instrumentation is aging and
deteriorating and that this has been the situation for some time... When I
sit on NSF study sections, the average age of the equipment to be replace
is 15 years (remember 1980? Reagan became President...)

The "bottom-line boy"s who now feel that they should run things
"efficiently" have cut spending on scientific infrastructure year after
year. Now, like the collapse of the school system in California, 20 years
after Prop. 13 "got gov't off their backs", the Universities are desparate
and are trying any old trick to keep afloat: tricks that, I may add, are
often heartily endorsed by the (other) for-profit enterpreneurs who now
hope to benefit from using university equipment and expertise that "they
have been paying for through their taxes for years"....

Everyone wants to cut the pie so that they get a little more and, it seems
to me, that this discussion is more about cupidity than anything else.
Ethics is an important topic, but, like wrapping oneself in the flag, there
is often a temptation to simplify it for the sake of making debating
points.

Not all things are well understood just by looking at the "Market Price".
After all, Chuck often contributes to our discussions in a helpful manner
but surely he doesn't suppose that his email connection fee actually covers
the cost of sending his thoughts to the hundreds of members of this list!

Jim Pawley

***************NEW ADDRESS**************
Prof. James B Pawley, Ph. 608-263-3147
Room 1235, Engineering Research Building, FAX 608-265-5315
1500 Johnson Dr. Madison, Wisconsin, 53706.
JPAWLEY-at-MACC.WISC.EDU






From: AWBlackwoo-at-aol.com
Date: Fri, 23 Jun 1995 14:38:16 -0400
Subject: SEM on Fluoridated Plastics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to Alex Titkov's posting on this subject, there is no "secret" to
imaging this sort of feature except to keep on trying! A couple of tricks
ought to help, however.

First, forget about doing any low magnification imaging on the same sample,
cut away literally all of the material that isn't in the picture and silver
paint everything that isn't in the picture (you are welcome to guess which
brand I prefer). If the sample is not completely grounded, you don't have a
chance of success, no matter what instrument you are using.

Second, you may have some success with putting down a very thin layer of
carbon before gold coating. It's no cure, but it sometimes helps.

Third, cut away more material. If you need a gold layer as thick as the
features you're trying to image, you've lost half of the battle before you
started. When it's the inside of a tube, I generally find that the "real"
problem is that I'm trying to image down in the groove of a semicircular
section; you really only want a small segment of the circle for best results.

Finally, work quickly. The less time the sample spends under the beam, the
greater the chance of success. One of the perverse things about our
instruments is that we seem to see much more beam damage during the rapid
scanning while we frame the picture and focus, and much less during the slow
scan exposure of the photographic emulsion. You might be able to get away
with focusing on one area and then obtaining the micrograph from another by
cranking the stage over, but this generally gives adequate focus only at low
magnification, in my experience, at least.

Good luck with a tough assignment.

Andy Blackwood

Andrew W. Blackwood, Ph.D.
Structure Probe, Inc.
63 Unquowa Road
Fairfield, CT 06430-5015
Ph: 1 203 254 0000
FAX: 1 203 254 2262
e-mail: AWBlackwoo-at-aol.com





From: Michael OKeefe :      Michael_OKeefe-at-macmail7.lbl.gov
Date: 23 Jun 1995 12:17:46 U
Subject: Re: HREM Shareware

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1408212583.67569-at-macmail7.lbl.gov}

Reply to: RE} } HREM Shareware

L.D. Marks wrote --

} The purpose of my email was not an attempt to kill either
} EMMPDL or MASSL but completely different. The idea is to collect
} primarily X-based routines for HREM and other miscellaneous such as
} Semper code and libraries. I believe that there has been a disturbing
} trend over the last ten years for heavy-duty programs for HREM to
} be written on Government funding then sold commercialy. These } programs
} should be much easier and cheaper than they are (} $5000 in many cases).
} I intend to donate NUMIS to this (at least the multislice and imaging
} parts) as soon as I can rewrite for X-windows and thereby escape a
} licensing agreement.
} L.D. Marks

------------------------------------------------------

For a HREM image simulation program that was "written on Government funding" and so is (and has always been) available free, see --
http://ncem.lbl.gov/NCEM/computer/software.html

BTW, this program is not available on the Zaluzec EMMPDL because of the requirement that all programs there include source code. I
have insisted that NCEM institute a policy of not releasing source code after my sad experience with my simulation code SHRLI. Over
the years SHRLI was written with various government funding support (CSIRO in Australia, NSF in the U.S., and SERC in the U.K.) and
was sent out free until my boss in Cambridge found out that one of the individuals who had requested SHRLI was now advertising and
selling it under a different name (he had made a small change -- he had taken the four main programs and combined two of them to
produce a suite of three). Of course, then my boss insisted that we charge for SHRLI and use the income to augment our meager
travel funds.

Also BTW, I am a little worried at the sound of your expressed intent to "donate NUMIS . . . as soon as I can rewrite for X-windows
and thereby escape a licensing agreement". Some people might think that rewriting code to escape a contractural agreement is
getting close to piracy . . . .

Michael A. O'Keefe
Acting Head, NCEM






From: David Henriks :      73531.1344-at-compuserve.com
Date: 23 Jun 95 15:36:50 EDT
Subject: RF Interscience address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can anyone provide me with the address, TEL and FAX for RF Interscience?

Thank you!

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com





From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Fri, 23 Jun 1995 14:50:25 -0500
Subject: Re: HREM Shareware

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rewriting code simply to change it in a minor fashion is certainly not appropriate, and
probably not legal. Just to clarify....I am talking about major changes in about 50-80%
of the current code of NUMIS (which, except for Public Domain Material) I wrote.




From: Dwight Beebe :      beebed-at-ERE.UMontreal.CA
Date: Fri, 23 Jun 1995 15:38:52 -0400 (EDT)
Subject: TEM: KMnO4 fixation of plant cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings!
I'm in kind of a bind. Does anyone have a protocol that they
would be willing to post directly to me regarding Mollenhauer's method of
KMnO4 fixation for plant cells? I'm interested only in a drastic method
for examining plant cell membranes and their continuity and I understand
Mollenhauer's protocol worked very well. Most texts and literature I
have direct access to consider this protocol outdated or too specialized
and so don't say much in practical terms about the how-to's.
Many thanks.

Dwight Beebe
Institut de recherche en biologie vegetale beebed-at-ere.umontreal.ca
Dept. de sciences biologiques Voice:514-872-4563
Universite de Montreal FAX:514-872-8496
4101, rue Sherbrooke est
Montreal, PQ H1X 2B2 Canada





From: Beverly E Maleeff :      Beverly_E_Maleeff%notes-at-sb.com
Date: 23 Jun 95 18:10:43 EDT
Subject: De-wrinkling thick sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a good method for removing very tiny wrinkles from half-micron
Epon-replacement sections?
My standard method, if chloroform stretching doesn't work, is to use sodium
ethoxide or methoxide etching. Unfortunately, in a sample I'm working on,
sodium ethoxide removed the cell components I'm studying (phospholipids) along
with the wrinkles. Are there any less destructive reagents out there?
TIA--

Bev Maleeff
SmithKline Beecham Pharmaceuticals
Toxicology-US, UE 0462
709 Swedeland Road
King of Prussia, PA 19406
610/270-7987
610/270-7202 fax
maleeffbe-at-sb.com or Beverly_E_Maleeff%Notes-at-sb.com





From: Louis M. Ross, Jr. :      geosclmr-at-showme.missouri.edu
Date: Sat, 24 Jun 1995 13:57:36 -0500
Subject: RE: Temp unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe microscopy geosclmr-at-showme.missouri.edu





From: goran.alsterborg-at-mailbox.swipnet.se (JEOL Service)
Date: Sat, 24 Jun 1995 21:41:56 +0000
Subject: Re: Digital SEM imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506241945.VAA04891-at-mailbox.swip.net}
X-Sender: m-13225-at-mailbox.swip.net (Unverified)
Mime-Version: 1.0
Content-Type: text/plain; charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

} Dear Greg,

} If you wish to passively digitize the image from your JEOL JSM 5300, this=
=20
can } be done by the JEOL SemAfore. This is a low cost passive digitizer=20
listening
} to the signal being sent to the photo CRT. It digitizes with the same=20
} resolution as the slow scan display on the photo CRT (typically 1900 x=
1400
} pixels) and displays/stores the image as a file in BMP format.=20

} After getting the image into your PC, you can correct greylevels, zoom,=20
count } etc and get the image on paper either as a playback to the photo CRT=
=20
or as a
} print on a LaserJet, video or dye sublimation printer.

} The digitizing board goes into a PC (IBM compatible 486 or faster with min=
8
} Mbyte RAM) and the program operates under MS Windows. It certainly works=
on
} the 5300 and also on almost all old or new JEOL SEM=B4s.'

} This unit is presently marketed only in Europe and Asia, so for more
} information, please contact us directly (preferably not by e-mail as we=
just
} have started testing the net and do not yet have any routines).

} Best regards from Goeran Alsterborg

} JEOL(Skandinaviska)AB
} Vegagatan 19 Fax: +46 - 8 - 29 16 47
} 172 34 Sundbyberg =20
} SWEDEN Phone: +46 - 8 - 28 28 00





From: Kym Smith :      KYMSMITH-at-immuno.imvs.sa.gov.au
Date: Mon, 26 Jun 1995 09:15:05 GMT+1030
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506252344.AA11941-at-huon.itd.adelaide.edu.au}

unsubscribe




From: D-PATTON-at-wpg.uwe.ac.uk
Date: Mon, 26 Jun 1995 10:14:38 +0000
Subject: LM MOLLUSCA CONCHIOLIN STAIN

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {sfee8eda.001-at-wpg.uwe.ac.uk}
X-Mailer: WordPerfect Office 4.0

Can anyone offer advice on the following problem?

The outer layer of molluscan shells (the periostracum) is a
proteinaceous sheet of conchiolin. After death the periostracum
rapidly decays. I wish to find a stain or reagent which will be
taken up by shells with a periostracum but rejected by shells
where the periostracum is lacking, (ie. a conchiolin-specific
stain or reagent.
Can anyone suggest a simple staining/reagent method for this??

From Karen Croker, C/O Dave Patton

(D-PATTON-at-uwe.ac.uk)
University of the West of England
Bristol
UK







From: KAKER-at-ctklj.ctk.si
Date: Mon, 26 Jun 1995 13:53:15 +0200 (WET-DST)
Subject: EDS intensity data

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

Does anyone have a EDS intensity data for testing standardless programs.
Up to now we are collected with our SEM Jeol JSM 35-CF and EDS Edax
9100/40, 205 spectra from different samples, under the well known
conditions (tilt angle, take-off angle, accelerating voltage, detector
parameters and sample composition).

Best regards,

Henrik Kaker

Metal d.o.o.
SEM/EDS Laboratory
Koroska c.14
62390 Ravne
SLOVENIA

Tel: +386-602-21-131/5562
Fax: +386-602-20-436
Internet E-mail: Kaker-at-ctklj.ctk.si





From: rybicka-at-acsu.buffalo.edu
Date: Mon, 26 Jun 1995 12:42:47 -0400
Subject: TEM - GLYCOGEN ULTRASTRUCTURE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, All Those Interested in Glycogen,

The existing confusion in glycogen staining, its disappearance
after use of uranyl acetate en bloc, and other problems of glycogen in TEM
are explained in recent short review printed in MICROSCOPY TODAY, October,
1994, entitled
"GLYCOGEN GFRANULES REVISITED".

Glycogen in the tissue contains covalently bound protein. The
event is well established in biochemistry, but still misinterpreted in EM.
Since this protein contains enzymes involved in glycogen metabolism, the
structures composed of glycogen-protein complex were considered as cellular
organelles and called GLYCOSOMES in 1968 ! So called "glycogen granules"
stainable with uranium and lead represent protein component (enzymes)
attached to glycogen in glycosomes. Glycogen in these organelles appears
as small (3 nm) particles which form (20-30nm) aggregates. These particles
remain invisible after routine staining with uranium and lead salts because
glycogen, as a polymer of glucose, does not form ionic bonds.

Apparent removal of "glycogen granules" by uranyl acetate, or other
acids, used before tissue dehydration, is in fact the removal of protein
component. The bond of protein to glycogen is sensitive to the change in pH
(this sensitivity was used in the traditional purification of glycogen by
the treatment with strong alkali or acids). The soluble protein separated
from glycosomes by uranyl acetate is presumably washed out during
dehydration. Glycogen is not fixed per se, but it is stabilized in
glycosomes due to the fixation of the bound protein. After removal of
protein by acidic treatment the structure of glycosomes is destroyed, and
small, 3 nm, glycogen particles floate freely in the cell forming large
irregular clumps. In routinely stained tissue these clumps appear as
unstained spots often misinterpreted as a poor plastic polymerization.
However, histochemical staining (Thiery technique) shows that these spots
are composed of the clumps of glycogen particles.

First TEM study of the subject was: RYBICKA, K.1979. Virchows
Archiv B. Cell Pathol. 30, 355-47. The article included prior references
showing how the misinterpretation of glycogen in TEM appeared. Recent
references are listed in the mentioned review in MICROSCOPY TODAY. If you
cannot get it, please, let me know, and I will write them.

Considering present status of biochemical knowledge about
glycosomes,
microscopic and molecular biology techniques seem to offer a good field for
further research.

Good luck,

Krystyna Kielan Rybicka
SUNY at Buffalo
Physiology / Neurobiology
Buffalo, NY


---
-------------------------------------------------------------------------
email address rybicka-at-acsu.buffalo.edu
phone number (lab) 716 829 3575
-------------------------------------------------------------------------





From: Edmund Glaser :      eglaser-at-umabnet.ab.umd.edu
Date: Mon, 26 Jun 1995 13:22:49 -0400 (EDT)
Subject: Re: Legal/Ethical Questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506261654.AA23256-at-cid.infosel.com.mx}

Usually I am a passive observer to the transactions of this group. However,
James Pawley's recent bilious response to Charles Garber's thoughtful remarks
made me sit up and take notice. I found no "diatribe" there, although that is
perhaps in the eye of the beholder. While I am sure that Dr. Garber can
defend himself, Prof. Pawley's shotgun response missed a glaring target right
in front of him and he doesn't seem to know it exists. He mentions the fact
that many outstanding universities require their faculty to support their
salaries from their consulting (translation: grants) 50% and sometimes more.
Indeed, this is so and indeed, it was not always so. In the more than 30
years that I've been in academia, this trend of reducing hard money salary
and converting faculty into entrepreneurs renting space in their schools has
been accelerating.

Some consequences of this are apparent:
(1) No grant, no position. This is also a not very subtle attack at tenure.
What does it mean to have tenure if you can't put food on the table?
(2) The "respected" universities are being subsidized by those institutions
who do pay hard money salaries. A small calculation shows that a researcher
who puts 50% of his salary on the grant is charging the granting institution
that amount plus fringe benefits (about 30% of salary) plus overhead
(minimum 40%) on the two for each year. There's then that much less for
everyone else on the pay line, and fewer on the pay line.

The pressure is intense for all academic research institutions to participate
in the market oriented industrial research game. They don't know whether to
treat their faculty as employees or as independent entrepreneurs, but certainly
not as pure academics. Result: a "beat up on the faculty" policy. Make them
support themselves even it drags them into creative financing to help them
survive.

What's this got to do with the original topic? Simply this. Don't blame
problems on how to stay both solvent and employed on industry and
entrepreneurs who know the rules they work under and are trying to make
a profit. The fault is university administrations who, having
succumbed to the lure of big (nonprofit) money in research, are making up
rules as they go and are playing corporate games on the
backs of their faculties. That's what gets the latter into the ethical/moral
and legal morass that started this whole thing off. Since the university
administrations got this problem going, let them solve it, if they
know how. One obvious answer is to have universities stop trying to be
money making machines and go back to what they're supposed to do, be
institutions of learning. I don't think this is likely to happen soon.


Edmund Glaser, D. Eng.
Dept. Physiol.
Univ. Md. School. Med.
Baltimore, MD 21201 USA
Ph: (410) 706-5041
Fax: (410) 706-8341





From: Michael Cammer :      cammer-at-aecom.yu.edu
Date: Mon, 26 Jun 1995 15:06:13 -0400 (EDT)
Subject: vibration table need

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


currently, we have floating vibration tables for our light microscopes.
We need one or two more and were wondering of cheaper alternatives, such
as tabletop units.

Any suggestions of vendors would be helpful.

Thank you--

Michael Cammer
cammer-at-aecom.yu.edu






From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Mon, 26 Jun 1995 14:23:40 -0500
Subject: Ge Chemical Thinning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a reasonable recipe for chemical thinning of Ge
which does not use Br (which has become a safety issue).

Thanks




From: Mike Bench :      bench-at-cems.umn.edu
Date: Mon, 26 Jun 1995 15:13:31 -0500
Subject: Re: Ge Chemical Thinning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In message {199506261923.AA18751-at-apollo.numis.nwu.edu} L.D.Marks writes:
} Does anyone have a reasonable recipe for chemical thinning of Ge
} which does not use Br (which has become a safety issue).
}
} Thanks

Laurie,
The method I have used (it has been about 5 years since I have done this) for
both Ge and Si, which is also a safety concern, is a 9 to 1 solution of HNO3 and
HF at, I think, room temperature. I suspended dimpled samples into a beaker of
the solution using platinum tipped tweezers. I can't remember if I stirred the
solution during thinnig or not. Notwithstanding the safety concerns with HF, I
found it easier to work with than Br in methanol solutions (which I have used
for GaAs but not Ge). As I remember it the thinning is pretty rapid, with it
taking not more than a few minutes to perforate a 50 micron thick sample so it
is difficult to get very small holes and very small wedge angles. Otherwise the
sample quality was quite good.
Good luck,



Mike Bench
Dept. of Chem Eng & Mat. Sci
or CIE Microscopy Center
250 Amundson Hall Office: 612-625-8508
421 Washington Ave. S.E. Microscopy Center: 612-624-6590
Univ. of Minnesota Fax: 612-626-7246
MInneapolis, MN 55455





From: EvexAnalyt-at-aol.com
Date: Mon, 26 Jun 1995 17:19:11 -0400
Subject: Announce: Training Video for TN5500 available from Evex

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


News Flash

Evex Analytical Instruments
(908) 874-3800

Evex has just produced a training Video for the Tracor Northern 5500 System

Future Videos to be released
Tracor Norther 2000
Noran Voyager
PGT IMIX
Kevex Delta




From: JOHNA-at-SCI.WFEB.EDU
Date: Mon, 26 Jun 1995 17:50:17 -0400 (EDT)
Subject: Re: TEM:diamond resharpening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Also not to run down any companies; I have been quite pleased with
MicroStar's resharpening service on my old DuPonts.

John

___________________________________________________
| |
| John G. Aghajanian, Ph.D. |
| Worcester Foundation for Experimental Biology |
| 222 Maple Avenue |
| Shrewsbury, MA 01545-2737 |
| |
| Tel: 508 842-8921 ext. 147, 161 |
| Fax: 598 842-9632 |
| JOHNA-at-sci.wfeb.edu |
| |
|_________________________________________________|





From: masur-at-msvax.mssm.edu
Date: Mon, 26 Jun 1995 19:56:45 -0500
Subject: Wild M40 Tissue Culture Microscope camera adapter parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


During a recent move, we have lost the trinocular assembly to attach a
Wild-modified Polaroid Land film adapter (which we still have)

The missing part(s)
#7567 phomicrographic attachment camera MKa2, fitting #3526 phototube
#3526 beamsplitter phototube


We will be delighted to purchase or barter for these parts

Lilly Ossowski, PhD
212-241-3194

or

Sandra K. Masur Box 1183
fax: 212-289-5945 Mount Sinai School of Medicine
phone: 212-241-6544 or 0089 NY NY 10029-6574
e-mail:masur-at-msvax.mssm.edu









From: David Henriks :      73531.1344-at-compuserve.com
Date: 27 Jun 95 12:22:35 EDT
Subject: Correction to Jet Thinning Germanium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The recipe I gave in my last message was for jet sectioning of polycrystalline
germanium. While the solution is the same, the parameters for jet thinning
single crystal germanium are as follows:

The solution is designated BK-1 and in short is:
500ml methanol
100ml butyl cellosolve
90ml H2SO4
30ml HF
Temperature: -50C
Jet height: 4.5mm
Flow Rate: Medium
Volts: 40
Current: 20mA

I'm sorry for any confusion.

Best regards-

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com





From: Christine A Brantner :      cab-at-csd.uwm.edu
Date: Tue, 27 Jun 1995 12:57:39 -0500 (CDT)
Subject: LR White/Immunostaining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi

I am a graduate student and I am trying to immunolocalize a protein to the
intracytoplasmic membranes of a bacteria.

Question?

Has anyone had experience infiltrating and embedding in LR White following
incomplete dehydration? Is 70% ETOH suffient? Does this improve the
retention of antigenicity for immunogold labeling? Does this protocol
require a modification of the polymerization schedule (time and/or
temperature)?

Thanks in advance. Any thoughts and comments would be helpful.

Chris Brantner

University of Wisconsin-Milwaukee
PO Box 413
Milwaukee, WI 53201

cab-at-csd.uwm.edu





From: GIDDINGS THOMAS H :      giddings-at-horton.Colorado.EDU
Date: Tue, 27 Jun 1995 09:18:31 -0600
Subject: ion milling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I'm looking for advice on performance testing of an ion milling
device. A group here at Univ of Colorado is using a Gatan PIMS to
mill titanium shadowed S-layers from Sulfolobus. Of late, they have
not been able to reproduce the high resolution pattern in milled
samples. We have evaluated the S-layers by TEM of negative stained
and shadowed preps and by checking the diffraction patterns of both,
and we have not detected any problems with either the biological
sample or the titanium. We would like to determine whether the PIMS
is performing properly. Is there a standard sample that can be used
to evaluate ion milling? Or, are there tests that can be performed?
We are obviously novices in the field; any advice appreciated.

Thanks,

Tom Giddings
giddings-at-horton.colorado.edu





From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Tue, 27 Jun 1995 07:42:27 -0400 (EDT)
Subject: RE: vibration table need

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Mon, 26 Jun 1995 15:06:13 -0400 (EDT),
Michael Cammer {cammer-at-aecom.yu.edu} writes:

} currently, we have floating vibration tables for our light microscopes.
} We need one or two more and were wondering of cheaper alternatives, such
} as tabletop units.Any suggestions of vendors would be helpful.
----
Partial list of table-top vibration isolation platform vendors & their Tel. Nos.:

Technical Maufacturing Corporation (TMC) 800-542-9725
Kinetic systems 800-992-2884
Barry Controls 617-787-1555
Newport Corp 714-253-1680
Herzan Vibration isolation systems (relatively economical models available)
714-859-7409

Good Luck.

*************************************************************************
M.V. Parthasarathy
Professor of Plant Biology & Director, EM Facility
Section of Plant Biology
228 Plant Science Building
Cornell University, Ithaca, NY 14850
Telephone: (607) 255-1734
Fax: (607) 255-5407
E-mail: mvp2-at-cornell.edu
***********************************************************************




From: wise-at-vaxa.cis.uwosh.edu
Date: Tue, 27 Jun 1995 13:11:30 +0000
Subject: Negative envelopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a source for white negative envelopes for TEM (3.5x4.5
inches) or SEM (4x5")? The only ones I can find are glassine or manila.

Bob Wise
Biology Department
UW Oshkosh
(414) 424-3404
wise-at-vaxa.cis.uwosh.edu






From: Goldmarker-at-aol.com
Date: Wed, 28 Jun 1995 06:23:43 -0400
Subject: Phalloidin Gold?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a source of phalloidin gold - 15 or 20nm?

Regards,

Donald P. Cox, Ph.D.
GOLDMARK BIOLOGICALS
437 Lock St
Phillipsburg, NJ 08865
(908) 859-2631
(908) 859-2875-FAX
goldmarker-at-aol.com




From: Clint E. Young :      clintey-at-unixg.ubc.ca
Date: Tue, 27 Jun 1995 23:46:46 -0700 (PDT)
Subject: Quantitative Immunohistochemistry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To fellow colleagues,

I am currently involved in trying to perform quantitative
immunohistochemistry using image analysis. Does anyone know of a
discussion group or mailing list on this topic?

Looking forward to your reply.

Clint Young
Department of Psychiatry
University of British Columbia






From: James R. Stets :      stetsjr-at-ttown.apci.com
Date: Wed, 28 Jun 1995 07:42:08 -0400 (EDT)
Subject: White negative envelopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bob Wise asked about a source for white envelopes for TEM and SEM
negatives. In the past, we have obtained them through Ladd Research
Industries, Inc.

Ladd Research Industries, Inc.
P.O. Box 1005
Burlington, VT 05402
802-658-4961

I think they are still in business; my catalog is a few years old.

Jim Stets
Air Products and Chemicals, Inc.
Allentown, PA
"I speak only for myself"




From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Tue, 27 Jun 1995 19:42:12 -0600
Subject: EM:micrographs for textbook

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v01510100ac166307bbff-at-[131.230.97.53]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We are revising a textbook on biological electron microscopy (SEM and TEM)
and are looking for interesting electron micrographs of biological
specimens. Especially we are interested in images of viruses, microwave
fixed specimens as compared to conventionally fixed specimens, plants, TEM
of fungi, insects and human norman tissue. Contact me directly if you have
such micrographs. Thank you in advance.

#############################################################################
John J. Bozzola, Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: Katherine.S.Connolly-at-Dartmouth.EDU (Katherine S. Connolly)
Date: 28 Jun 95 13:14:42 EDT
Subject: negative envelopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {17232611-at-dancer.Dartmouth.EDU}

Crimson Tech
325 Vassar St.
Cambridge, Mass 02139

Kate Connolly
Dartmouth Med School




From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Wed, 28 Jun 1995 10:19:19 -0700
Subject: Re: Negative envelopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d01ac1739f11315-at-[129.82.126.28]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Does anyone have a source for white negative envelopes for TEM (3.5x4.5
} inches) or SEM (4x5")? The only ones I can find are glassine or manila.
}

I know that glassine envelopes are not OK for archival storage of
negatives, and manila are probably not either. Check some of your old
negatives that have been stored in them, and you'll likely find some
discoloration/darkening under the glued seam.

We now use transparent polyethylene negative storage envelopes from
Polysciences.

Cat. No. Neg. size latest price
07233 3.25in X 4in $29.70
07235 4in X 5in $32.30

In addition to providing better protection, you can look at the negatives
on a light table without taking them out of their envelopes.

--John






From: Lauri J. Pelliniemi :      ljpelmi-at-utu.fi
Date: Tue, 27 Jun 1995 08:00:51 -0900 (PDT)
Subject: Re: glycogen fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear listreaders:
Glycogen is an interesting object, which is prone to cause problems
especially for unexperienced beginners. I have tried to explain parts of
the issue in: Pelliniemi et al.: Glycogen accumulations in
differentiating mesonephric ducts and tubuli in male human embryos. 1983
Anatomy and Embryology 168:445-453. A comparative table of glycogen
behaviour in different procedures is included.
Yours, Lauri
-----------------------------------------------------------------------------
I Dr. Lauri J. Pelliniemi I Telephone +358-21-633 7312 I
I Associate Professor I or +358-21-633 7209 I
I Laboratory of Electron Microscopy I I
I University of Turku I Telefax +358-21-633 7380 I
I Kiinamyllynkatu 10 I Internet LJPELMI-at-UTU.FI I
I FIN-20520 Turku I Bitnet/EARN LJPELMI-at-FIRIEN.BITNET I
I FINLAND, Europe I I
-----------------------------------------------------------------------------





From: David Henriks :      73531.1344-at-compuserve.com
Date: 27 Jun 95 12:14:45 EDT
Subject: Ge Chemical Thinning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Laurie-

You may want to refer to the paper by Bernie Kestel "A Jet Electropolishing
solution for Silicon, Germanium, Tantalum, Niobium and Tungsten-Rhenium"
Ultramicroscopy 9 (1982) 379-384. If you prefer, I can send you a copy . It is
only 5 pages long so I could even FAX it if you need it fast. Just e-mail me or
give us a call and ask for a copy of Tech paper #5.

The solution is designated BK-1 and in short is:
500ml methanol
100ml butyl cellosolve
90ml H2SO4
30ml HF
Temperature: -60C
Jet height: 6.5mm
Flow Rate: Fast
Volts: 150
Current: 95mA

Kestel writes:

"The small amount of HF in the BK-1 removes thin oxides well, yet creates only a
moderate hazard. The butyl cellosolve widens the voltage plateau for optimal
polishing and improves the finish obtained across precipitates and grain
boundries "
This work was done with the Model 550 Single Vertical jet electropolisher.
Voltage, temp etc. will have to be adjusted if you are using a twin jet system.

I hope this helps!

Best regards-

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com





From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Wed, 28 Jun 1995 13:45:42 -0500
Subject: Re: Ge Chemical Thinning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


With thanks to various people for recipes on Ge thinning, I still have
one caveat and one question.
The caveat is that we are going to use the Ge samples in our UHV microscope.
As a consequence large hydrocarbon molecules such as butyl cellosolve are
the last thing that we want -- it may take forever to remove these to obtain
clean surfaces.
The question concerns what chemical for what crystallographic orientations?
We have found experimentaly that the standard Si polish works well for Ge
(001) but fails for Ge (111). Does anyone know of recipes for Ge (111) ?




From: John W Heckman :      heckman-at-pilot.msu.edu
Date: Tue, 27 Jun 1995 20:12:58 -0400 (EDT)
Subject: Re: Color Correction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506280012.AA15582-at-pilot01.cl.msu.edu}

Damian:

The color correction you depends on the film type you are using as
well as the output freqs. of the light source. For Kodak Ektachrome and
Kodachrome 64's 30M +2/3 stop is recommended for average fluorescent light.
Exact correction depends on just what type of tubes are in the light (and their
age too I'd suspect). For more info, might check Kodak Pub E-77 or the latest
update thereof (If you are still in a jam I can Fax the table).

cheers,
John Heckman
heckman-at-pilot.msu.edu
} } } Good
Morning Everyone }
} I know this is supposed to be a microscopy forum, BUT, I'm in a jam and need
} to know
} what CC filters to use to correct for standard fluorescent lamps using
} daylight film. I had
} a sheet that listed the different filtration packs for different lamps but
} can't find it. I have
} a wide selection of large CC filters, know that the correction is mostly a
} magenta filter but
} also remember that other CC colors were used. No time to hunt down a 72 mm
} FLD filter!
}
} Many thanks to anyone who can help.
}
} Damian
} neuberd-at-roundlake.baxter.com
}





From: JOHNA-at-SCI.WFEB.EDU
Date: Wed, 28 Jun 1995 09:42:39 -0400 (EDT)
Subject: Re: LR White/Immunostaining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Chris,

Yes, LR White is great stuff for IMC. DO NOT use Osmium as per their
instruction sheet. Dehydration through 70% EtOH is sufficient and you can
dehydrate through absolute EtOH. You probably should test both to see how
your antigen reacts. If you don't have complete polymerization after
overnite and 50-55 C, leave the blocks at 60 C for a day (I try to avoid
the latter, however.). Good luck.

John

___________________________________________________
| |
| John G. Aghajanian, Ph.D. |
| Worcester Foundation for Experimental Biology |
| 222 Maple Avenue |
| Shrewsbury, MA 01545-2737 |
| |
| Tel: 508 842-8921 ext. 147, 161 |
| Fax: 598 842-9632 |
| JOHNA-at-sci.wfeb.edu |
| |
|_________________________________________________|





From: braunfeld-at-msg.ucsf.edu
Date: Wed, 28 Jun 1995 12:15:44 -0700
Subject: EPMA Beam deflection

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506281056.AA52602-at-pukrs7.puk.ac.za}

Has anyone out there in the cryo world heard of an Italian company that
makes cold stages which are supposed to be a less expensive alternative
to either gatan or oxford??? I've heard mention of one but do not know
who they are ar how to contact them. If anybody knows something please
let me know. thank you.




From: Brian J. Zande :      bz0c+-at-andrew.cmu.edu
Date: Wed, 28 Jun 1995 16:30:57 -0400 (EDT)
Subject: unsuscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {IjwPk1C00iV9Q6VAIf-at-andrew.cmu.edu}


please unsuscribe




From: DRStadden:R_D:Armstrong
Date: 6-28-95 3:30pm
Subject: OM Polymer Staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: MICROSCO:CENTRAL
cc: DRStadden:R_D:Armstrong
Subj: OM Polymer Staining
------------------------------------------------------------------

I'm wondering if there may be a good course one might recommend in the
subject area of staining methods for optical microscopy of polymers. I
know the subject has been dealt with in various texts, but can't say
that I know of any formal educational opportunities in this area.
Also, if it included stains useful to SEM of polymer composites, that
would be a plus. My experience is with PLM and SEM of flooring and
ceiling products, gasketing and felt products, and rubber insulation and
textile mill products, as well as components thereof. Thanks in advance
for any suggestions.

Dave Stadden
DRStadde+aCENTRAL%ARMSTRONG-at-MCIMAIL.COM

Ph: 717-396-5109
FAX: 717-396-5865





From: Tamara A Howard :      howie+-at-pitt.edu
Date: Wed, 28 Jun 1995 15:52:51 -0400 (EDT)
Subject: TEM tech position open

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9506282122.AA07137-at-worldlink.worldlink.com}


TEM Technician Position Open

The Department of Pathology's EM Research Facility is seeking qualified
applicants for an EM technician position; a background in cell biology is
preferred. The applicant should have experience in specimen preparation
for TEM, particularly in ultrathin sectioning a wide variety of
specimens. Experience in EM autoradiography and immunocytochemistry are
not required but would be extremely helpful. The person in this position
will be responsible for handling research EM projects for academic and
clinical pathologists, post-docs, residents, and fellows.
Salary commensurate with experience.
The University of Pittsburgh is an Equal Opportunity Employer
Contact:
Fang He
S-704 Scaife
Cell/Molecular Pathology
3550 Terrace St.
U. Pittsburgh School of Medicine
Pittsburgh, PA 15261
(412) 648-9467
fan-at-med.pitt.edu




From: dennbarr-at-emn.com (Dennis B. Barr)
Date: Tue, 27 Jun 1995 08:49:34 -0400
Subject: RE Vibration table need

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I recently saw a description of a vibration-free platform for a table top from
Terra Universal, Inc., 700 N. Harbor Blvd., Anaheim, CA 92805.

Our lab once used a couple of table-top units from another manufacturer (who is
no longer in business). The pneumatic isolators in those units effectively
reduced vibration to acceptable levels.

Among other possible sources there are Kinetic Systems, Inc., 20 Arboretum
Road, P.O. box 414, Boston, MA 02131, and Technical Manufacturing Corporation,
15 Centennial Drive, Peabody, MA 01960.
Dr. Dennis B. Barr
Eastman Chemical Company
Microscopy and Morphology Research Laboratory
Kingsport, TN 37662-5150

Phone: 615/229-2188
E-mail: dennbarr-at-emn.com
FAX: 615/229-4558






From: kennedy-at-nsi.edu (grace kennedy)
Date: Wed, 28 Jun 1995 15:28:01 -0800
Subject: Lr White IM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: grace-at-198.147.151.5
Message-Id: {ac17943f02021004e631-at-[198.147.151.19]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

In addition to John Aghajanian's recommendations, I would suggest the
following two papers: "An Osmium-free Method of Epon Embedment That
Preserves both Untrastructure and Antigenicity for Post-embedding
Immunocytochemistry", by Phen, K.D., Rustioni, A., and Weinberg, R.J., J.
Histochem.Cytochem. 42, 1995 This method should adapt nicely to LR White.
Sorry I don't have page numbers. And, "An Enhanced Method for
Post-embedding Immunocytochemical Staining Which Preserves Cell Membranes",
by Berryman, M.A., and Rodewald, R.D., J. Histochem. Cytochem. 38 (2)
159-170, 1990. Also would adapt nicely to LR White material. Grace
Kennedy






From: Steven Bagley :      SBAGLEY-at-fs2.scg.man.ac.uk
Date: Tue, 27 Jun 1995 16:35:00 GMT+1
Subject: vibration tables

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} currently, we have floating vibration tables for our light microscopes.
} We need one or two more and were wondering of cheaper alternatives, such
} as tabletop units.
}
Hi Michael,

A idea I have used for a confocal microscope is using a concrete
table. On top of the table a slab on slate or granite - from your
friendly local stone mason. Sandwiched between the two are four heavy
duty inner tubes with special valves which do not allow air out when
the tubes are pumped up.

In the sandwich layer you may also wish to put four wooden blocks
thus the table can be leveled by using these blocks as a reference.
This may sound a 'wacky' set up but this table has been in use four
five years with no problems yet.

Hope this has helped

regards,


Steve Bagley

__________________________________________________________________

The Confocal Microscope Facility
Cells, Immunology and Development Voice: 0161 275 6771
University of Manchester Fax: 0161 275 3915

http://gonzo.sci.man.ac.uk/~sbagley/

If you tied buttered toast to the back of a cat and
dropped it from a height, what would happen ?
___________________________________________________________________




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Wed, 28 Jun 1995 11:24:02 +1100
Subject: Critical Point Dryer/Dye Sub Printers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: st004718-at-brandywine.otago.ac.nz
Message-Id: {v01510102ac164d8539df-at-[139.80.120.172]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Thanks very much for all those who were kind enough to write regarding
their experiences with their dye sublimation printers. There are very few
of these in NZ and its reassuring to get feedback from users.

On a different note, we are seriously looking at buying a Balzers CPD 030
critical point dryer and again would be grateful to hear from anyone who
has one as to whether there is anything we should be aware of with regards
to its use (or for that matter its accessories). We had our fingers burnt
three years ago when the university bought two new critical point dryers
(NOT Balzers) which turned out to be extremely unreliable and had to have
extensive and expensive modifications done to make them useable, even so
they have never really been satisfactory. We want to avoid a repeat of that
experience at all costs as we are very lucky to get funding again for a
replacement. I'm sure the Balzers instrument is in a different (better)
league altogether but if anyone has any good or bad stories with the CPD
030, I'd like to hear.
Regards and thanks,

Richard

Richard Easingwood
South Campus Electron Microscope Unit
Otago Medical School
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

"The southernmost electron microscope unit in the world"






From: chandler
Date: Wednesday, June 28, 1995 10:19AM
Subject: Re: Negative envelopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Does anyone have a source for white negative envelopes for TEM (3.5x4.5
} inches) or SEM (4x5")? The only ones I can find are glassine or manila.
}

I know that glassine envelopes are not OK for archival storage of
negatives, and manila are probably not either. Check some of your old
negatives that have been stored in them, and you'll likely find some
discoloration/darkening under the glued seam.

We now use transparent polyethylene negative storage envelopes from
Polysciences.

Cat. No. Neg. size latest price
07233 3.25in X 4in $29.70
07235 4in X 5in $32.30

In addition to providing better protection, you can look at the negatives
on a light table without taking them out of their envelopes.

--John






From: shirley.turner-at-NIST.GOV (Shirley Turner)
Date: Wed, 28 Jun 1995 13:14:35 -0400 (EDT)
Subject: ED: NIST Workshop on Crystallographic Databases

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Members of the electron diffraction community that are interested in using
crystallographic databases in materials characterization are invited to
attend the 'NIST Workshop on Crystallographic Databases' to be held on
August 29-30, 1995 at the National Institute of Standards and Technology. A
main goal of the Workshop is to foster interactions between users and
providers of crystallographic databases and between the communities that use
the different databases. Three sessions - each with seven or more
presentations- will be held and are summarized below:

(1) Formal Data Activities; Chair: David Watson, Cambridge
This session will focus on discussion of present activities and future
activities associated with the following data centers:

The Metals Database
Inorganic Crystal Structure Database (ICSD)
The Protein Data Bank
NDB- Nucleic Acid Database
The Cambridge Structural Database
The Powder Diffraction File
NIST Crystal and Electron Diffraction Data Center

(2) Scientific Uses of the Databases; Chair: Carolyn Brock, Univ. of Kentucky
This session will focus on using crystallographic databases in analysis, in
the prediction of materials properties and in the design of new chemicals,
pharmaceuticals and materials.

(3) Data Transfer: Ensuring state-of-the-art technology; Chair: Brian
McMahon, IUCr
This session will focus on issues related to data transfer such as:

(1) data exchange standards (CIF, etc.),
(2) the role of journals in the evaluation of published data;
(3) data exchange between the journals and crystallographic data
centers;
(4) computerized modes of data dissemination.

Following the presentations, two discussion sessions will focus on Barriers
to the Use of Crystallographic Data and on Partnerships for the Future.
Workshop proceedings will be published in a special issue of the NIST
Journal of Research. There is no registration fee and attendance will be
limited in order to hold the Workshop to an efficient and practical size.
Please contact us if you would like further information or registration forms.

Dr. Vicky Lynn Karen karen-at-tiber.nist.gov
Dr. Alan D. Mighell mighela-at-tiber.nist.gov
NIST Crystal and Electron Tel: 301-975-6255
Diffraction Data Center Fax: 301-975-2128
Materials Building, Room A215
Nat'l. Inst. of Stds. & Tech.
Gaithersburg, Maryland 20899





From: GVKM07A-at-prodigy.com (DR CHARLES A GARBER)
Date: Tue, 27 Jun 1995 20:28:05 EDT
Subject: Different legal issue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

I have received in my mail box so far a total of 31 "private" replies
to my postings of a week ago in response to John Wheatley's original
posting. This subject clearly strikes the nerves of more than a few
who follow this listserver.

One especially interesting issue was raised in several of these private
notes and it had to do with the "legal" side of the discussion but from
a different perspective: Can a client sue a university faculty member
(or the university itself) if the client thinks the university made a
mistake and caused the client economic loss?

I suspect that this will come as a surprise to some (many?) but the
answer is very definitely "yes!".

The US court system has made it very clear that when someone holds
themselves out as offering a professional service, be it a faculty
member offering electron microscopy services or a neurosurgeon doing
brain surgery, then the public has a right to expect a minimum level of
proficiency and competence and when the public perceives they did not
get that minimum level, then under the present laws and court system,
the injured party has a right to seek some relief through that court
system, and to recover damages they might have suffered as a
consequence. This kind of exposure is called "professional risk" and
one protects their assets by purchasing professional liability
insurance.

I have not verified the statistics, but I have been told by one person
who has looked at this that someone "consulting" is more likely to get
caught up in some kind of a professional liability action than he is
in his life time to run down and kill someone with his car. Yet would a
responsible person ever dream of driving their car for even one minute
without automobile insurance?

Among the independent analytical and testing laboratory community,
there is unanimous recognition of the need to have professional
liability insurance coverage. No reputable and financially responsible
laboratory would be without such coverage. Laboratories do get sued on
a regular basis. I am also including microscopy laboratories in that
statement. Such actions might not make CNN or Peter Jennings, but they
do occur on a regular basis. One can spend $100,000 without any effort
at all on a typical case in federal court, and even if the judge
ultimately dismisses the "consultant" from the case, there is still
nevertheless the obligation for the payment of the legal fees and
share of the court costs.

No university, public or private, has been granted any kind of
immunity from such suits. So the point is that in order to do this
"commercial" work, in order to bring in this extra money, despite the
potential for disruption to legitimate academic programs and also to
the commercial marketplace, this activity puts the entire university at
risk in ways that are never even considered. And for a public
institution, the real losers are the taxpayers, who now have to carry
the buden or risk of the folly of the commercial activities being
conducted out of the university's laboratories.

Chuck

======================================================
Charles A. Garber, Ph. D.
President
SPI SUPPLIES Div. of Structure Probe, Inc.
PO BOX 656
West Chester, PA 19381-0656 USA

Ph: 1-(610)-436-5400
1-(800)-2424-SPI

FAX: 1-(610)-436-5755

e-mail: GVKM07A-at-prodigy.com
SpiSupp-at-aol.com [SPI Customer Service e-mail box]
======================================================





From: Clint E. Young :      clintey-at-unixg.ubc.ca
Date: Wed, 28 Jun 1995 17:17:40 -0700 (PDT)
Subject: Agar bound to slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To fellow colleagues,

Can anyone suggest how I can make chunks of agar adhere to slides? I've
tried using poly-lysine coated slides and saline coated slides, but both
require that the agar be DRIED onto the slides--something I want to avoid.

Looking forward to your reply.

Clint Young
Department of Psychiatry
University of British Columbia




From: fr.Bob Roberts i.n.d.c. :      bobrob-at-cam.org
Date: Sat, 24 Jun 1995 08:52:46 -0400
Subject: Re: (Fwd) A little English humor for the week-end (long_

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-isnet.is.wfu.edu *
**************************************************************

---------- Forwarded message ----------

Hi everybody
Hi everybody
Summer is here and some of you might be travelling.
And this can happen to you! ;-)

} *********************************************************************
}
}
} Subj: Hotel soap
} Date: Wed, Jun 21, 1995 12:13 AM EDT
}
} RE: Hotel soap
}
} Attached is some correspondence which actually occurred between a
} London hotel's staff and one of its guests. The London hotel
} involved submitted this to the Sunday Times. No name was mentioned.
}
} ******************************************************************************
} Dear Maid,
} Please do not leave any more of those little bars of soap in my
} bathroom since I have brought my own bath-sized Dial. Please remove
} the six unopened little bars from the shelf under the medicine chest
} and another three in the shower soap dish. They are in my way. Thank
} you,
} S. Berman
} ----------------------------------------------------------------------
} Dear Room 635,
} I am not your regular maid. She will be back tomorrow, Thursday,
} from her day off. I took the 3 hotel soaps out of the shower soap
} dish as you requested. The 6 bars on your shelf I took out of your
} way and put on top of your Kleenex dispenser in case you should
} change your mind. This leaves only the 3 bars I left today which my
} instructions from the management is to leave 3 soaps daily.
} I hope this is satisfactory.
} Kathy, Relief Maid
} ----------------------------------------------------------------------
} Dear Maid -- I hope you are my regular maid.
} Apparently Kathy did not tell you about my note to her concerning the
} little bars of soap. When I got back to my room this evening I found
} you had added 3 little Camays to the shelf under my medicine cabinet.
} I am going to be here in the hotel for two weeks and have brought my
} own bath-size Dial so I won't need those 6 little Camays which are on
} the shelf. They are in my way when shaving, brushing teeth, etc.
} Please remove them.
} S. Berman
} ----------------------------------------------------------------------
} Dear Mr. Berman,
} My day off was last Wed. so the relief maid left 3 hotel soaps which
} we are instructed by the management. I took the 6 soaps which were
} in your way on the shelf and put them in the soap dish where your
} Dial was. I put the Dial in the medicine cabinet for your
} convenience. I didn't remove the 3 complimentary soaps which are
} always placed inside the medicine cabinet for all new check-ins and
} which you did not object to when you checked in last Monday. Please
} let me know if I can of further assistance.
} Your regular maid,
} Dotty
} ----------------------------------------------------------------------
}
}
}
} Dear Mr. Berman,
} The assistant manager, Mr. Kensedder, informed me this A.M. that you
} called him last evening and said you were unhappy with your maid
} service. I have assigned a new girl to your room. I hope you will
} accept my apologies for any past inconvenience. If you have any
} future complaints please contact me so I can give it my personal
} attention. Call extension 1108 between 8AM and 5PM. Thank you.
} Elaine Carmen
} Housekeeper
}
} ----------------------------------------------------------------------
} Dear Miss Carmen,
} It is impossible to contact you by phone since I leave the hotel for
} business at 745 AM and don't get back before 530 or 6PM. That's the
} reason I called Mr. Kensedder last night. You were already off duty.
} I only asked Mr. Kensedder if he could do anything about those little
} bars of soap. The new maid you assigned me must have thought I was a
} new check-in today, since she left another 3 bars of hotel soap in my
} medicine cabinet along with her regular delivery of 3 bars on the
} bath-room shelf. In just 5 days here I have accumulated 24 little
} bars of soap. Why are you doing this to me?
} S. Berman
}
} ----------------------------------------------------------------------
} Dear Mr. Berman,
} Your maid, Kathy, has been instructed to stop delivering soap to your
} room and remove the extra soaps. If I can be of further assistance,
} please call extension 1108 between 8AM and 5PM. Thank you,
} Elaine Carmen,
} Housekeeper
}
} ----------------------------------------------------------------------
} Dear Mr. Kensedder,
} My bath-size Dial is missing. Every bar of soap was taken from my
} room including my own bath-size Dial. I came in late last night and
} had to call the bellhop to bring me 4 little Cashmere Bouquets.
} S. Berman
}
} ----------------------------------------------------------------------
} Dear Mr. Berman,
} I have informed our housekeeper, Elaine Carmen, of your soap problem.
} I cannot understand why there was no soap in your room since our
} maids are instructed to leave 3 bars of soap each time they service a
} room. The situation will be rectified immediately. Please accept my
} apologies for the inconvenience.
} Martin L. Kensedder
} Assistant Manager
}
} ----------------------------------------------------------------------
}
}
} Dear Mrs. Carmen,
} Who the hell left 54 little bars of Camay in my room? I came in last
} night and found 54 little bars of soap. I don't want 54 little bars
} of Camay. I want my one damn bar of bath-size Dial. Do you realize
} I have 54 bars of soap in here. All I want is my bath size Dial.
} Please give me back my bath-size Dial.
} S. Berman
}
} ----------------------------------------------------------------------
} Dear Mr. Berman,
} You complained of too much soap in your room so I had them removed.
} Then you complained to Mr. Kensedder that all your soap was missing
} so I personally returned them. The 24 Camays which had been taken
} and the 3 Camays you are supposed to receive daily (sic). I don't
} know anything about the 4 Cashmere Bouquets. Obviously your maid,
} Kathy, did not know I had returned your soaps so she also brought 24
} Camays plus the 3 daily Camays. I don't know where you got the idea
} this hotel issues bath-size Dial. I was able to locate some
} bath-size Ivory which I left in your room.
} Elaine Carmen
} Housekeeper
} ----------------------------------------------------------------------
} Dear Mrs. Carmen,
} Just a short note to bring you up-to-date on my latest soap
} inventory.
}
} As of today I possess:
}
} - On shelf under medicine cabinet - 18 Camay in 4 stacks of
} 4 and 1 stack of 2.
} - On Kleenex dispenser - 11 Camay in 2 stacks of 4 and 1
} stack of 3.
} - On bedroom dresser - 1 stack of 3 Cashmere Bouquet, 1
} stack of 4 hotel-size Ivory, and 8 Camay in 2 stacks of 4.
} - Inside medicine cabinet - 14 Camay in 3 stacks of 4 and 1
} stack of 2.
} - In shower soap dish - 6 Camay, very moist.
} - On northeast corner of tub - 1 Cashmere Bouquet, slightly used.
} - On northwest corner of tub - 6 Camays in 2 stacks of 3.
}
} Please ask Kathy when she services my room to make sure the stacks
} are neatly piled and dusted. Also, please advise her that stacks of
} more than 4 have a tendency to tip. May I suggest that my bedroom
} window sill is not in use and will make an excellent spot for future
} soap deliveries. One more item, I have purchased another bar of
} bath-sized Dial which I am keeping in the hotel vault in order to
} avoid further misunderstandings.
} S. Berman
}
} ---------------------------------------------------------
} "Water is composed of two gins, Oxygin and Hydrogin. Oxygin is pure
} gin. Hydrogin is gin and water."
}

Soon Y. Choi

"What is history but a fable agreed upon?" - Napoleon B.



| /\_/\
| (- -)
\_-------=\"/= email: bobrob-at-cam.org
| ______ / URL: http://www.cam.org/~bobrob/
|| ||
|| || God Bless you one and all fr. Bob+ {8-))
^^ ^^














From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Thu, 29 Jun 1995 07:58:04 +0100 (BST)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I reply to the request about Italian Companies who might make inexpensive
cold stages, you might try contacting Prof. Ugo Valdre, Department of
Physics, University of Bologna, Bologna, Italy. He has made cold stages
in the past and may well know of the contacts in Italy

\Good luck in your search

Patrick Echlin
Director, Multi-Imaging Centre
University of Cambridge
On Wed, 28 Jun 1995
braunfeld-at-msg.ucsf.edu wrote:

} Has anyone out there in the cryo world heard of an Italian company that
} makes cold stages which are supposed to be a less expensive alternative
} to either gatan or oxford??? I've heard mention of one but do not know
} who they are ar how to contact them. If anybody knows something please
} let me know. thank you.
}




From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Thu, 29 Jun 1995 08:07:10 +0100 (BST)
Subject: Re: (Fwd) A little English humor for the week-end (long_ (fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The person who is so concerned about hotel soap surely must have been
staying at Fawlty Towers !!

Patrick Echlin




From: Romuald Wroblewski onkpat :      Romuald.Wroblewski-at-onkpat.ki.se
Date: Thu, 29 Jun 1995 09:43:13 +0200 (METDST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Unsubscribe (for the summer)
I will be back in September.
Thanks for the humor in latest messages from London hotel. I loved it
although it is not directly forum for this type of messages but from time
to time it is OK.

-------------------------
Romuald Wroblewski, Ph.D.
Associate Professor
Department of Pathology
Karolinska Institute
Stockholm, Sweden
voice:+46-8-7293597
-------------------------





From: digigene-at-pavilion.co.uk (Dr L Van der Pant)
Date: Thu, 29 Jun 1995 08:55:56 +0100
Subject: DIGITAL TEM CCD OR FILM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,
I have been listening with a great deal of interest the debate about the
pros and cons of digital imaging. The company that I work for manufacture
cooled CCD cameras and interfaces for TEM's. In my experience the following
is true.
1. For biological samples at low magnification you lose image definition,
i.e. it becomes impossible to define a fine structure (membrane typically) at a
precise mag. i.e. at 2K you can see it at 1K you can't.Therefore it is
possible for
some simple teste on a given sample to determine when exactly the camera
system "loses it". Work on this basis
with a 1,000x1,000 pixel camera at the low end magnification would seem to
indicate that 2,000x2,000 pixel cameras will make the whole thing a lot more
viable.
2. Digital cameras give much better results than conventional TV.
3. YAG scintillant provides a consistant image in terms of uniformity of
illumination with no low structure (granularity) present even with high dynamic
imaging to 12-bits.

The other preferences dicussed really depend on whether your primary
interest is "the image" or results from the image i.e. gold labelleing
analysis. On this basis most people
tend to live with the CCD and forego the joys of the dark room.

As far as the future goes 2Kx2K cameras are here, but are very expensive.
Cost effective options will be available with 2Kx2K sensors, but probably
with column defects. But if they are reproducible then they can generally be
removed in the computer.


Leslie Vanderpant
DIGITAL PIXEL LIMITED
PO BOX 625
BRIGHTON BN1 5JT
UNITED KINGDOM

t) 00 44 1273 502176
f) 00 44 1273 502176





From: zvert-at-r1.atki.kfki.hu (Zofia Vertesy)
Date: Thu, 29 Jun 1995 10:03:46 +0200
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html








From: Rolf Odselius :      Rolf.Odselius-at-emu.lu.se
Date: Thu, 29 Jun 1995 14:27:40 +0100
Subject: Re: Critical Point Dryer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Richard Easingwood wrote:
} ...................we are seriously looking at buying a Balzers CPD 030
} critical point dryer and again would be grateful to hear from anyone who
} has one as to whether there is anything we should be aware of with regards
} to its use (or for that matter its accessories). We had our fingers burnt
} three years ago when the university bought two new critical point dryers
} (NOT Balzers) which turned out to be extremely unreliable and had to have
} extensive and expensive modifications done to make them useable, even so
} they have never really been satisfactory. We want to avoid a repeat of that
} experience at all costs as we are very lucky to get funding again for a
} replacement. I'm sure the Balzers instrument is in a different (better)
} league altogether but if anyone has any good or bad stories with the CPD
} 030, I'd like to hear.

We have used the CPD 030 for several years in a multi-user
laboratory (many students). I have found it extremely reliable. The
only problem we have had was small Teflon pieces detaching from the
magnet valves, thus jamming the gas outlet aperture. This problem
is solved since long. Also the accessories are well made and very
suitable for their purpose (in contrary to the accessories of the
earlier models).

Rolf

========================================================================
Rolf Odselius, PhD, Ass Prof
Electron Microscopy Unit, University Hospital
S-221 85 Lund, Sweden
Phone: +46 46 171075 Fax: +46 46 172975 Mobile phone: +46 10 6705655
Pager (Minicall): +46 740 288992 E-mail: Rolf.Odselius-at-emu.lu.se
URL: http://www.wblab.lu.se/medfak/medinst/emu/
Educational Secretary, SCANDEM URL: http//www.ldc.lu.se/~scandem/




From: Chris Frethem (CBN) :      frethem-at-lenti.med.umn.edu
Date: Thu, 29 Jun 1995 08:09:02 -0500 (CDT)
Subject: Negative sleeves/envelopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I agree with recommendation for use of polyethylene negative sleeves
for long term storage and convenience when using light box. Check with
your local poly bag manufacturing companies; I recently ordered 5000
4.125" X 5.5" 1.5 mil sleeves for Polaroid T-55 negatives locally
for $20.37 / 1000. They could probably make other sizes as well but I
didn't ask. Some companies don't want to be bothered with custom sizing
or small orders, so shop around. I don't want to advertise for this
particular supplier, but if you contact me directly I will give you
name and address.

Chris





=======================================================================
Chris Frethem (612)624-4652 (voice)
Cell Biology & Neuroanatomy (612)624-8118 (FAX)
U of MN, Minneapolis e-mail: frethem-at-lenti.med.umn.edu







From: EMLAB-at-opus.mco.edu
Date: Thu, 29 Jun 1995 09:32:55 -0500 (EST)
Subject: Re: Agar bound to slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Clint,

Try using glue. Loctite Tissue Adhesive is used all the time for gluing
tissues to aluminium blocks for vibrtome sectioning. It is available thru
Ted Pella.

Good Luck,
Ed Calomeni
Medical College of Ohio
Toledo, OH
emlab-at-opus.mco.edu




From: Walt Bobrowski :      bobroww-at-aa.wl.com
Date: Thu, 29 Jun 1995 10:28:33 -0400 (EDT)
Subject: 'Reduced' Osmium Query

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta PETVAX.MUAS; Relayed; Thu, 29 Jun 1995 10:28:33 -0400
Mr-Received: by mta PETVAX; Relayed; Thu, 29 Jun 1995 10:28:35 -0400
Mr-Received: by mta SRVR01; Relayed; Thu, 29 Jun 1995 10:29:57 -0400
Disclose-Recipients: prohibited

Can someone please differentiate the use of potassium ferrocyanide-reduced OsO4
for immunoelectron microscopy and plain 'ol OsO4 in lower concentration?

Thanks in advance,

Walt Bobrowski
Parke-Davis Research
Ann Arbor, MI






From: rcrang-at-pop.life.uiuc.edu (Richard Crang)
Date: Thu, 29 Jun 1995 10:07:27 -0500
Subject: Containing Unicells for Freeze-Substitution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are preparing plant (beet root) protoplasts for transmission electron
microscopy studies and would like to have some of them prepared by means of
freeze substitution. They are in an osmotically adjusted buffer medium as a
cell suspension and, while we can centrifuge them (gently) to condense them
and even freeze them in chilled propane, they will disperse during the
substitution process, and particularly upon warming. Thus, they will be
lost upon the exchange of fluids. We are using a Reichert CS-Auto FS
system. Any suggestions how the material (or any unicellular preparations
for that matter) can be handled to contain them during freeze substitution?
Maintaining the cells on agar is not possible due to the need to keep them
in an osmotic medium prior to freezing. Suggestions greatly appreciated.



**************************
Richard F. E. Crang
Professor of Plant Biology
University of Illinois
(217) 244-3143
**************************





From: Yuan Liang Chen :      cheny-at-mcnc.org
Date: Thu, 29 Jun 1995 12:29:29 -0400
Subject: Re: Gen Info: Starting the next millenia

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dr. Zaluzec,
My subscription to AAEM.AMC.ANL.GOV was undeliverable yesterday.
I has been off the internet for 6 months and don't know what
happened. Also I can not reach BBS. Please tell me what to do.
Thanks.
YL_CHEN-at-PNL.GOV




From: bafpjec-at-uxa.ecn.bgu.edu (Joyce Craig)
Date: Thu, 29 Jun 1995 11:35:18 -0500
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please take me off the microscopy listserver
Joyce Craig
Biology Department
Chicago State University
9501 King Drive
Chicago, IL 60628





From: emlab-at-ucsco.ucsc.edu (Jon Krupp)
Date: Thu, 29 Jun 1995 10:05:52 -0700
Subject: Nikon Microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506291704.KAA01886-at-ucsco.ucsc.edu}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

The following is posted as a courtesy for a colleague not familiar with
this list.

You may recall my request for help identifying a Nikon microscope earlier
this month. As you can see, we were able to find out a lot, due in part to
helpful readers of this list. Thanks to all. This list and all of you are a
great resource for me and many others at UC Santa Cruz.
***************************************************************************


Here is the more complete data on the Nikon X-Y microscope we
communicated about earlier this month:

6x6 Nikon Measuremaster, precision measuring X-Y binocular
microscope system for transmitted and reflected light, X-Y stage allows for
6" movement in each direction, separate Fiber-lite high intensity fiber
optic illumination system (series-180), Nikon digital X-Y counter (CM-65),
manual stage drives, but can be motorized, five turret rotating objective
head with Nikon x2.5 & x10 objectives, Vickers x20 objective, and Zeiss x25
& x100 objectives, tower allows for microscope to be focused up about 6" to
10" above stage (for large samples); microscope was not much used and is in
excellent condition.

Asking Price: $6000

If you're interested in it and have a funding source, let me know.

best wishes,
Othmar Tobisch
408-459-2777
otobisch-at-earthsci.ucsc.edu

*************************************************************************

Jonathan Krupp
Microscopy and Imaging Lab
University of California
Santa Cruz, CA 95064
(408) 459-2477
emlab-at-ucsco.ucsc.edu






From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 29 Jun 1995 16:48:47 -0400
Subject: Re-IntersciAddr

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1407677874.79134-at-mse.engin.umich.edu}

Subject: Time: 4:44 PM
OFFICE MEMO Re:IntersciAddr Date: 6/29/95

I believe that Interscience is a division of John Wiley & Sons, 605 Third
Ave., NY, NY, 10158. Tel: 212-850-6000





From: wise-at-vaxa.cis.uwosh.edu
Date: Thu, 29 Jun 1995 16:17:11 +0000
Subject: Re-IntersciAddr

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To all,

Occassionally, I have run across references to saphire or vitreous
carbon knives for ultramicrotomy but I have never seen any for sale. Does
anyone know if such knives were ever (are) commercially available?

Bob Wise


Robert R. Wise, PhD
Director, UWO Electron Microscope Facility
Department of Biology
University of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404
(414) 424-1101 fax
wise-at-vaxa.cis.uwosh.edu






From: Clint E. Young :      clintey-at-unixg.ubc.ca
Date: Thu, 29 Jun 1995 23:08:50 -0700 (PDT)
Subject: Quantitative Image Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To fellow collegues

There does not seem to be a forum (mailing list, discussion group, etc..)
singularly devoted to the subject of quantitative image analysis (Q-IA).
I would, therefore, like to create a mailing list of sorts.

First, let me explain what my lab is doing.

In my lab, quantitative image analysis (Q-IA) is used in combination with
immunohistochemistry and Western Blotting. The purpose of Q-IA, in the
context of our lab work, is to capture images of immunostained tissue and
immunoblots in order to determine the amount of staining and, therefore,
the amount of antigen present.

Q-IA of immunostained tissue can also be used to determine:

A. MACROSCOPIC FIELD
1. Number of cells stained
2. Size of cells
3. Distribution of cells

B. MICROSCOPIC FIELD
1. Amount of antigen present
2. Distribution of antigen
3. Colocalization of different antigens (confocal)

Q-IA of immunoblots, meanwhile, is used primarily for B.1-2.

To perform this type of analysis one needs peripherals (digital camera
and/or flatbed scanner), hardware (PC or Macintosh computer), and
software (NIH Image, Adobe Photoshop, etc...).

Obviously, Q-IA can be performed using many different setups depending on
the combination of peripherals-hardware-software used.

I would like, therefore, to make a list of those interested in this topic
and will circulate this list among the participants.

Please include the following:

1. E-Mail address
2. Computer
3. Image Analysis Software
4. Type of analysis you are CURRENTLY ABLE to perform
5. Question/Comment


----------
The following is an example of what I would send:

clintey-at-unixg.ubc.ca
Macintosh
NIH Image
IHC: A(1,2,3) + B(1,2,3)
WB: B(1,2)
?: I am taking images of Western Blots using reflective light; I've tried
making the WB transparent but the process causes the staining to run or
streak. Does anyone know which method--reflective or transmitted
light--works best for quantifying Western Blots?

----------

Please E-Mail me directly since this message will be sent to various
mailing lists.

I will send participants the compiled list on Monday July 3.

Looking forward to your reply.

Clint Young
Department of Psychiatry
University of British Columbia





From: rms-at-vax.ox.ac.uk
Date: Fri, 30 Jun 1995 11:15:19 +0100
Subject: J. Microsc. Abstracts. Delete message if not interested!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

JTERLET-at-CEMMA.ADELAIDE.EDU.AU, microtoday-at-aol.com, ROSS.MACKENZIE-at-OX.AC.UK,
MICROSCOPY-at-AAEM.AMC.ANL.GOV
Message-ID: {00992A63.3FDF0AA6.23-at-vax.ox.ac.uk}

JOURNAL OF MICROSCOPY ABSTRACTS
DECEMBER 1994 - FEBRUARY 1995

FOR MORE DETAILS ABOUT THE JOURNAL OF MICROSCOPY, CONTACT RMS-at-VAX.OX.AC.UK


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
181-187.

In situ analysis of microbial consortia in activated sludge
using fluorescently-labelled, rRNA-targeted oligonucleotide
probes and scanning confocal laser microscopy

M. Wagner, B. Assmus, A. Hartmann, P. Hutzler & R. Amann
Lehrstuhl fur Mikrobiologie, Technische Universitat Munchen,
Arcisstrasse 21, D-80290 Munchen, Germany


SUMMARY

Activated sludge flocs are complex consortia of various micro-
organisms. The community structures of samples taken from
municipal sewage treatment plants were characterized using
fluorescently labelled, 16S and 23S rRNA-targeted
oligonucleotide probes in combination with confocal scanning
laser microscopy (CSLM). In comparison with conventional
epifluorescence microscopy, CSLM considerably improved the
capability to visualize directly the spatial distribution of
defined bacterial populations inside the sludge flocs.
Analyses could be performed at high resolution undisturbed by
problems such as autofluorescence or limited spatial
resolution in thick samples. In addition, CSLM was used to
analyse some structural properties of paraformaldehyde-fixed
activated sludge flocs, such as floc size and homogeneity.
Typical floc sizes were found to be in the range between 5 and
50 micrometre. Whereas most of the flocs were completely
colonized by bacteria, there were also examples of flocs
containing gas bubbles or particles in the interior.


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
188-194.

Scanning interference and confocal microscopy

R. Juskaitis & T. Wilson, Department of Engineering Science,
University of Oxford, Parks Road, Oxford, OX1 3PJ, U.K.


SUMMARY

The form of the interference term image in scanning confocal
and scanning conventional interference microscopes is
identical in all respects including optical sectioning. This
observation is used to obtain confocal images and surface
profiles from conventional scanning interference microscope
images.


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
195-203.

Time- and wavelength-resolved spectroscopy in two-photon
excited fluorescence microscopy

S. Andersson-Engels, I. Rokahr & J. Carlsson
Department of Physics, Lund Institute of Technology, PO Box
118, S-221 00 Lund, Sweden


SUMMARY

Two-photon excited fluorescence spectroscopy has been
performed at a microscopic scale in combination with normal,
white light microscopy. This gave simultaneously a spectral
resolution of 20nm and a temporal resolution of 20ps, from a
volume element less than 5 micrometre in all three dimensions.
The sample was excited with light from a continuously mode-
locked Ti:sapphire laser that was focused on the sample in a
fluorescence microscope. A polychromator and streak-camera
were used for detection. The method has been used on tissue,
plant and paper samples. It has also been demonstrated how
substances naturally occurring in the samples can be
identified from their spectroscopic properties and the spatial
distribution of these substances can be observed.



Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
204-210.

Intracellular localization of the antitumour drug adriamycin
in living cultured cells: a confocal microscopy study

S. Meschini, A. Molinari, A. Calcabrini, G. Citro & G. Arancia
Department of Ultrastructures, Istituto Superiore di Sanita,
Viale Regina Elena 299, 00161 Rome, Italy


SUMMARY

The intracellular distribution of the anthracyclinic
antibiotic adriamycin in living cultured cells has been
investigated by confocal microscopy.
In human melanoma cells (M14), adriamycin was localized
inside the nuclei. When adriamycin-treated M14 cells were
allowed to recover in a drug-free medium, a complete efflux of
the drug from the nucleus was revealed. In recovered cells, a
weakly fluorescent signal was observed in the perinuclear
region. When M14 cells were recovered in a medium containing
colcemid, a microtubule depolymerizing agent, the drug
transport from the nucleus to the cell periphery appeared to
be inhibited, suggesting that the microtubule network is
strongly involved in drug transport mechanisms. In multidrug-
resistant (MDR) cells the intracellular location of adriamycin
was shown to be noticeably different from that of the parental
wild-type cells. In particular, in resistant human breast
carcinoma cells (MCF-7), adriamycin appeared to be exclusively
located within the cytoplasm, whereas the nuclei were shown to
be completely negative. When adriamycin treatment was
performed in association with MDR revertants, such as
Lonidamine (inhibitor of the energy metabolism) or verapamil
(inhibitor of the P-glycoprotein efflux pump), a marked
enhancement of the cytoplasmic signal was observed in
resistant cells. Under these conditions, adriamycin appeared
concentrated in the perinuclear region, but the nuclei were
still negative. Confocal microscopy proved to be a useful
method for the study of the intracellular transport of
fluorescent substances, such as anthracyclinic antibiotics,
and for the investigation of the multidrug resistance
phenomenon in tumour cells.



Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
211-221.

A versatile tilting device for fluorescence microscopes

J. Bradl, M. Hausmann, B. Schneider, B. Rinke & C. Cremer
Institut fur Angewandte Physik, Albert-Uberle Strasse 3-5,
D-69120 Heidelberg, Germany


SUMMARY

A tilting device for biological specimens (rotation angle up
to 2 pi) especially fluorescence-labelled cell nuclei, was
developed. It consists of a quartz glass capillary and a
mounting adaptor for the microscope stage. The applicability
of the device was tested for several epifluorescence and
confocal scanning laser microscopes. The axis of rotation is
perpendicular to the optical axis of the microscope. The
capillary can be tilted around its axis at any desired angle
or in equiangular steps. This can be done manually or by
remote control using a stepping motor.
The three-dimensional (3-D) image-forming properties of
the capillary system were experimentally examined using an
inverse confocal scanning laser microscope. The results were
compared with measurements obtained from the same microscope
with the standard stage for plane slides with cover glasses.
The measured point spread function suggested that, in spite of
aberration effects, the optical arrangement used allows a gain
in the 3-D resolution by tilting the object.
A low-cost, fully-automated 3-D imaging system was built
on the basis of a conventional epifluorescence microscope with
a cooled black-and-white CCD camera. The system was operated
by a personal computer. The online visualization ('movie') of
rotating objects indicates the feasibility of the system.



Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
222-225.

Continuous wave excitation two-photon fluorescence microscopy

P. E. Hanninen, E. Soini & S. W. Hell
Department of Medical Physics, University of Turku, Center for
Biotechnology, Tykistokatu 6, FIN-20521 Turku, Finland


SUMMARY

Two-photon excitation fluorescence imaging is feasible with
continuous wave lasers. Images of biological specimens are
obtained by employing photon counting in conjunction with an
increasing recording time. The approach allows two-photon
three-dimensional imaging of fluorescently-labelled specimens
with inexpensive lasers.


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
226-230.

Refractive-index-induced aberrations in two-photon confocal
fluorescence microscopy

H. Jacobsen, P. E. Hanninen, E. Soini & S. W. Hell
Department of Medical Physics, University of Turku, Center for
Biotechnology, Tykistokatu 6, FIN-20521 Turku, Finland


SUMMARY

The effect of refractive index mismatch on the image quality
in two-photon confocal fluorescence microscopy is investigated
by experiment and numerical calculations. The results show a
strong decrease in the image brightness using high-aperture
objectives when the image plane is moved deeper into the
sample. When exciting at 740nm and recording the fluorescence
around 460nm in a glycerol-mounted sample using a lens of a
numerical aperture of 1.4 (oil immersion), a 25% decrease in
the intensity is observed at a depth of 9 micrometre. In an
aqueous sample, the same decrease is observed at a depth of 3
micrometre. By reducing the numerical aperture to 1.0, the
intensity decrease can be avoided at the expense of the
overall resolution and signal intensity. The experiments are
compared with the predictions of a theory that takes into
account the vectorial character of light and the refraction of
the wavefronts according to Fermat's principle. Advice is
given concerning how the effects can be taken into account in
practice.


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
231-237.

The tetrahedral tip as a probe for scanning near-field optical
microscopy at 30nm resolution

U. C. Fischer, J. Koglin & H. Fuchs
Westfalische Wilhelms Universitat, Physikalisches Institut,
Wilhelm-Klemm-Strasse 10, 48149 Munster, Germany


SUMMARY

The tetrahedral tip is introduced as a new type of probe for
scanning near-field optical microscopy (SNOM). Probe
fabrication, its integration into a scheme of an inverted
photon scanning tunnelling microscope and imaging at 30nm
resolution are shown. A purely optical signal is used for
feedback control of the distance of the scanning tip to the
sample, thus avoiding a convolution of the SNOM image with
other simultaneous imaging modes such as force microscopy. The
advantages of this probe seem to be a very high efficiency and
its potential for SNOM at high lateral resolution below 30nm.



Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
238-244.

Studies of porphyrin containing specimens using an optical
spectrometer connected to a confocal scanning laser microscope

O. Trepte, I. Rokahr, S. Andersson-Engels & K. Carlsson
Physics IV, The Royal Institute of Tech, S-100 44 Stockholm,
Sweden


SUMMARY

A spectrometer has been developed for use with a confocal
scanning laser microscope. With this unit, spectral
information from a single point or a user-defined region
within the microscope specimen can be recorded. A glass prism
is used to disperse the spectral components of the recorded
light over a linear CCD photodiode array with 256 elements. A
regulated cooling unit keeps the detector at 277K, thereby
allowing integration times of up to 60s. The spectral
resolving power ranges from 350 at 400nm to 100 at 700nm.
Since the entrance aperture of the spectrometer has the same
size as the detector pinhole used during normal confocal
scanning, the three-dimensional spatial resolution is
equivalent to that of normal confocal scanning. Light from the
specimen is deflected to the spectrometer by a solenoid
controlled mirror, allowing fast and easy switching between
normal confocal scanning and spectrometer readings.
With this equipment, studies of rodent liver specimens
containing porphyrins have been made. The subcellular
localization is of interest for the mechanisms of photodynamic
therapy (PDT) of malignant tumours. Spectroscopic detection is
necessary to distinguish the porphyrin signal from other
fluorescent components in the specimen. Two different
substances were administered to the tissue, Photofrin, a
haematoporphyrin derivative (HPD) and delta-amino levulinic
acid (ALA), a precursor to photoporphyrin IX and haem in the
haem cycle. Both are substances under clinical trials for PDT
of malignant tumours. Following administration of these
compounds to the tissue, the potent photosensitizer and
fluorescent compound photofrin, or protoporphyrin IX,
respectively, is accumulated. For our study Wistar/Furth rats
were injected either with Photofrin or with ALA 3-5h before
they were killed. The organs were removed directly after and
snap-frozen in carbon dioxide ice with isopentane. No further
staining or fixation procedures were adopted.


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
245-253.

Modelling of inclined and curved surfaces in the reflection
scanning acoustic microscope

W. Weise, P. Zinin & S. Boseck
Institute for Materials Science and, Structure Research,
Physics Department, University of Bremen, 28334 Bremen,
Germany


SUMMARY

An expression is derived for the output signal when an
inclined plane surface is imaged by the reflection scanning
acoustic microscope, which is modelled as a spherical
transducer. This expression is applied to model non-planar
surfaces. The accuracy of this approach is tested for
perfectly reflecting spherical surfaces. The influence of
inclination on V(z) curves is considered when Rayleigh waves
occur.


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
254-261.

Scanning force microscopy on live cultured cells: imaging and
force-versus-distance investigations

D. Ricci & M. Grattarola
Dipartimento di Ingegneria Biofisica, ed Elettronica,
Universita degli Studi di Genova, Via Opera Pia 11a, 16145
Genova, Italy


SUMMARY

Extensive measurements with the scanning force microscope
(SFM) on living cells in their native liquid environment are
described with the purpose of critically assessing the extent
of the interaction between the SFM tip and the (soft) cell
materials and the effect of such interaction on topographic
information. Images are obtained under various force
conditions and systematically correlated with force-versus-
distance curves. As a result, detailed indications about tip
indentation are given, thickness estimates deduced and
identification of submembranous cytoplasmic structures
suggested.


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
262-275.

In vivo analysis of angiogenesis and revascularization of
transplanted pancreatic islets using confocal microscopy

F. A. Merchant, S. J. Aggarwal, K. R. Diller & A. C. Bovik
Biomedical Engineering Research Program, ENS 612, University
of Texas, Austin 78712-1084, U.S.A.


SUMMARY

A technique to measure angiogenesis and revascularization in
pancreatic islets transplanted at the renal subcapsular site
in the rat has been developed. In vivo imaging of the
microcirculation of transplanted pancreatic islets was
conducted using a confocal scanning laser microscope (CSLM) to
achieve optical sectioning through the graft in order to
perform a computer reconstruction of the three-dimensional
neovascular morphology. Individual islets were harvested by
enzymatic digestion of excised pancreas from Fischer 344 rats.
Isolated islets were cultured for 24h, and approximately 300-
350 islets were transplanted at the renal subcapsular site of
the left kidney in an anaesthetized rat. Six to 14 days post-
transplantation, the animal was anaesthetized and prepared for
in vivo imaging of the microvasculature on a Zeiss LSM-10.
Optical contrast of the microvasculature was enhanced by the
administration of fluorescein-labelled dextran into the
circulating blood. The transplant site was identified and
serial sections were obtained through the vascular bed at
varying z-intervals. Complementary fluorescence video images
were also obtained via a silicon intensifier tube camera
mounted on the CSLM. At completion of the imaging procedure,
the kidney was returned into the body cavity, the area was
sutured and the animal was allowed to recuperate for
subsequent examinations. Image processing algorithms, such as
grey-level thresholding, median filtering, skeletonization and
template matching, were applied to compute the vessel density
and diameters and extrapolated to measure 3-D vessel lengths
and the tortuosity index of the neovasculature.



Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
276-280.

Optoelectronic detector probes for scanning near-field optical
microscopy

H. U. Danzebrink
Physikalisch-Technische Bundesanstalt, Bundesallee 100,
D-38116 Braunschweig, Germany


SUMMARY

A brief explanation of the optoelectronic probe concept and a
comparison between the implementation of passive waveguide
probes and optoelectronic probes in scanning near-field
optical microscopy (SNOM) is presented. The first probe
realizations using cleaved semiconductor crystals and the work
at present in progress using microfabricated Si pyramids are
described. These crystals with evaporated metal electrodes
forming a slit aperture with subwavelength dimensions work as
metal-semiconductor-metal photodetectors. Their optical
detection behaviour is investigated by measuring the intensity
distribution of a laser focal point. Measurements where the
external bias voltage is changed show that it is possible to
modify the detection behaviour of the device because of the
varying depletion widths. The last part of the article
describes a concept where pyramidal probes should function
simultaneously as senors for scanning force microscopy (SFM)
to measure topography and as optoelectronic probes for
scanning near-field optoelectronic microscopy (SNOEM).


Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
281-286.

Imaging in the far-red with electronic light microscopy:
requirements and limitations

C. Cullander
School of Pharmacy S926, University of California, San
Francisco, CA 94143-0446, U.S.A.


SUMMARY

The acquisition of simultaneous dual confocal images with red
and far-red light has both advantages (e.g. lower
autofluorescence) and limitations. An understanding of these
requisites is necessary to acquire high-quality images and to
avoid the misinterpretation of experimental data. The poor
detection of far-red light mandates a high optical transfer
efficiency for the system, thus the transmittance of the
objective lens and its axial and lateral chromatic aberration
in the far-red are important factors for consideration. This
technical note is an attempt to 'demystify' the process of
filter set design for confocal microscopy by discussing the
considerations that went into the construction of a filter set
for use with the reagents cyanine 3.18 (Cy3) and cyanine 5.18
(Cy5), and thus to encourage users to look beyond the
multipurpose designs available commercially. The 568-nm laser
line exciting Cy3 is at its emission maximum, which limits the
collectable Cy3 fluorescence. High-transmission optical
filters with sharp bandpass cutoffs are thus desirable for
maximum light throughput. Light path mirror efficiency rapidly
degrades above 700nm, but the loss of this portion of the Cy5
emission spectrum is acceptable since the fluorophore is very
bright, and these very long wavelengths are also likely to
introduce aberration. While resolution is decreased with far-
red light, there is also greater penetration and less
scattering, and it is thus possible to obtain high-quality
images from deeper within the specimen. Although only one make
and model of confocal microscope (the Bio-Rad MRC-600) is
considered, similar considerations pertain to the design of
filter sets for any confocal microscope that accommodates
user-installed filters.



Journal of Microscopy, Vol. 176, Pt. 3, December 1994, pp.
287-299.

Simultaneous confocal recording of multiple fluorescent labels
with improved channel separation

K. Carlsson, N. Aslund, K. Mossberg & J. Philip
Physics IV, The Royal Institute of Tech, S-100 44 Stockholm,
Sweden


SUMMARY

Confocal microscopes are often used to study specimens
labelled with fluorophores. A commonly used method for
simultaneous recording of the distribution of multiple
fluorophores is to divide the fluorescent light emitted by the
specimen into different wavelength regions using dichroic and
bandpass filters. These different wavelength regions are then
distributed to multiple detectors. However, fluorophores often
result in considerable cross-talk between channels. A new
technique, intensity-modulated multiple-beam scanning (IMS)
microfluorimetry, can be used to reduce this cross-talk
substantially.
The IMS technique is implemented with two laser beams of
different wavelengths, intensity-modulated at different
frequencies, which illuminate the specimen simultaneously. The
two laser wavelengths predominantly excite one fluorophore
each. Fluorescent light from the specimen is divided into two
wavelength regions (red and green) which are detected by two
photomultiplier tubes. The output signals from the
photomultiplier tubes are connected to lock-in amplifiers. The
effect of using modulated laser beams, in combination with
lock-in amplifiers, is strongly to reduce the cross-talk
between channels. The performance of the IMS technique using
various types of specimen is compared with the results
obtained using the conventional multi-detector design.




Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp. 1-6.

In vivo determination of fibril orientation in plant cell
walls with polarization CSLM

J.-P. Verbelen & D. Stickens
Department of Biology, University of Antwerp (UIA),
Universiteitsplein 1, B-2610 Wilrijk-Antwerpen, Belgium


SUMMARY

Congo Red fluorescence is used to detect cellulose in the wall
of plant cells. The orientation of the cellulose fibrils is
determined by using polarized light for excitation. The
absorption characteristics of Congo Red make this approach a
method of choice for applications with any standard confocal
scanning laser microscopy (CSLM). The semi-quantitative
character of CSLM observations combined with the non-toxicity
of the stain allow a very fast and reliable assessment of
cellulose orientation in the wall of living plant cells.



Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
7-17.

Three-dimensional reconstruction of the human renal glomerulus

K. Preston Jr, B. Joe, R. Siderits & J. Welling
Kensal Corporation, 5055 East Broadway (Suite C206), Tucson AZ
85711, U.S.A.


SUMMARY

The capillary bed of human renal glomerulus is one of the more
complex capillary structures in the human body. This paper
illustrates three-dimensional reconstruction of the capillary
bed from serial sections. It shows that, although traditional
methods of three-dimensional rendering by computer fail to
handle the complexities of the capillary structure, new
methods based on filtering using three-dimensional
mathematical morphology are capable of revealing previously
unseen details. This is done at the expense of eliminating
fine structure (small capillaries). An error analysis allows
the degree to which fine details are lost to be estimated.



Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
18-30.

Quantitative water mapping of cryosectioned cells by electron
energy loss spectroscopy

S. Q. Sun, S.-L. Shi, J. A. Hunt & R. D. Leapman
Health & Human Services, Public Health Service, National
Institutes of Health, Building 13 Room 3W13, Bethseda MD
20892, U.S.A.


SUMMARY

A direct technique based on electron energy-loss spectroscopy
(EELS) in the scanning transmission electron microscope (STEM)
has been developed to map subcellular distributions of water
in frozen-hydrated biological cryosections. Previously,
methods for water determination have been indirect, in that
they have required the cryosections to be dehydrated first.
The new approach makes use of spectrum-imaging, where EELS
data are collected in parallel at each pixel. Several
operations are required to process the spectra including:
subtraction of the detector dark current, deconvolution by the
detector point-spread function, removal of plural inelastic
scattering and correction for the support film. The resulting
single scattering distributions are fitted to standard
reference spectra at each pixel, and water content can be
determined from the fitting coefficients. Although the
darkfield or brightfield image from a hydrated cryosection
shows minimal structure, the processed EELS image reveals
strong contrast due to variation in water content. Reference
spectra have been recorded from the major biomolecules
(Protein, lipid, carbohydrate, nucleic acid) as well as from
vitrified water and crystalline ice. It has been found that
quantitative results can be obtained for the majority of
subcellular compartments by fitting only water and protein
reference spectra, and the accuracy of the method for these
compartments has been estimated as plus/minus 3.5%. With the
present instrumentation the maximum allowed dose of 2000e/nm2
limits the useful spatial resolution to around 80nm plus/minus
5% precision at a single pixel. By averaging pixel intensities
a value of 56.8% with a precision of plus/minus 2.0% has been
determined for the water content of liver mitochondria. The
water mapping technique may prove useful for applications to
cell physiology and pathophysiology.



Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
31-42.

Hybrid scanning transmission electron/scanning tunnelling
microscope system for the preparation and investigation of
biomolecules

H. F. Knapp, R. Wyss, R. Haring, C. Henn, R. Guckenberger & A.
Engel
Maurice E Muller Institut fur, Hochauflosende
Elektronenmikroskopie, Universitat Basel, Klingelbergstrasse
70, CH-4056 Basel, Switzerland


SUMMARY

A hybrid scanning transmission electron/scanning tunnelling
microscope vacuum system is introduced, which allows freeze
drying and metal coating of biological samples and their
simultaneous observation by scanning transmission electron
microscopy and scanning tunnelling microscopy (STM). Different
metal coatings and STM tips were analysed to obtain the
highest possible resolution for such a system. Bovine liver
catalase was used as a test sample and the STM results are
compared to a molecular scale model.


Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
43-52.

Backscattered electron imaging of the undersurface of
resin-embedded cells by field emission scanning electron
microscopy

R. G. Richards & I. Ap Gwynn
AO/ASIF Research Institute, Clavadelerstrasse, CH-7270 Davos
Platz, Switzerland


SUMMARY

In this study backscattered electron (BSE) imaging was used to
display cellular structures stained with heavy metals within
an unstained resin by atomic number contrast in successively
deeper layers. Balb/c3T3 fibroblasts were cultured on either
13mm discs of plastic Thermanox, commercially pure titanium or
steel. The cells were fixed, stained and embedded in resin and
the disc removed. The resin block containing the cells was
sputter coated and examined in a field-emission scanning
electron microscope. The technique allowed for the direct
visualization of the cell undersurface and immediately
overlying areas of cytoplasm through the surrounding embedding
resin, with good resolution and contrast to a significant
depth of about 2 micrometre, without the requirement for
cutting sections. The fixation protocol was optimized in order
to increase heavy metal staining for maximal backscattered
electron production. The operation of the microscope was
optimized to maximize the number of backscattered electrons
produced and to minimize the spot size. BSE images were
collected over a wide range of accelerating voltages (keV),
from low values to high values, to give 'sections' of
information from increasing depths within the sample. At 3-
4keV only structures a very short distance into the material
were observed, essentially areas of cell attachment to the
removed substrate. At higher accelerating voltages information
on cell morphology, including in particular stress fibres and
cell nuclei, where heavy metal were intensely bound, became
more evident. The technique allowed stepwise 'sectional'
information to be acquired. The technique should be useful for
studies on cell morphology, cycle and adhesion with greater
resolution than can be obtained with any light-microscope-
based system.



Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
53-67.

Tomographic reconstruction of the cross-sectional refractive
index distribution in semi-transparent birefringent fibres

T. C. Wedberg & W. C. Wedberg
Physics Department, University of Bergen, Allegt 55, N-5007
Bergen, Norway


SUMMARY

Optical diffraction tomography (ODT) is used to reconstruct
the complex refractive index distribution in cross-sections of
semi-transparent, birefringent fibres. The selected fibres
were polymer and animal fibres of either circular or non-
circular cross-section with average thicknesses in the range
8-110 micrometre. This choice of samples was made to
illustrate the imaging capabilities of ODT, and also to
demonstrate some potential applications of the technique. The
images representing the reconstructed refractive index
distributions have a spatial resolution of about 2 micrometre,
and show noticeable image contrast for refractive index
variations of about 0.001. The ODT reconstructions compare
well with refractive index information provided with the
samples, and with scanning electron micrographs of cross-
sections of the same fibre samples. From these results it
appears that ODT can be used to reconstruct the complex
refractive index distribution in cross-sections of semi-
transparent birefringent fibres.



Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
68-76.

Computer simulated high-resolution transmission electron
microscopy (HRTEM) in tourmaline

E. A. Ferrow
Avdelningen for Mineralogi och Petrologi, Geologiska
Institutionen, Lund Universitet, Solvegatan 13, 223 62 Lund,
Sweden


SUMMARY

The contrast distributions observed in high-resolution
transmission electron microscopy (HRTEM) images of tourmaline
depend on the types and magnitudes of the exchange components
present and on the degree of atom overlap along the direction
of observation. Furthermore, the fractional atomic coordinates
in tourmalines are valid only for the specific specimen
refined. These properties make the interpretation of
experimental HRTEM images of tourmaline using image simulation
if not impossible at least extremely difficult. A correct
interpretation of experimental HRTEM images of tourmaline is
possible provided the structural refinement data on the same
crystal are available. Nevertheless, it is possible to
interpret the experimental HRTEM images of tourmaline if the
composition of the structural model chosen during image
simulation approximates the composition of the specimen
studied by electron microscopy. A good control of the
composition of the specimen studied and an appropriate choice
of a structural model for image simulation are therefore as
important as properly controlling specimen thickness, specimen
tilt, beam tilt and objective lens defocus.



Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
77-84.

Confocal microscopy in the analysis of the etched nuclear
particle tracks in polymers

J. Jakes, P. Gais & H. Schraube
Institut fur Strahlenschutz, GSF-Neuherberg, Postfach 1129,
D-85758 Oberschleissheim, Germany


SUMMARY

The possibility of the morphometric analysis of etched tracks,
induced by protons and alpha particles in the organic polymer
allyl diglycol carbonate (CR-39), using the confocal scanning
laser microscope (CSLM), was studied. The detectors were
investigated in two groups of irradiation experiments, namely:
(a) irradiated with mono-energetic neutrons of energy 1.2MeV,
(b) exposed to the alpha radiation from 222Rn and its progeny.
Both groups were irradiated at normal incidence. Radiation-
induced latent tracks were electrochemically etched, and their
morphometric parameters were investigated in the reflection
mode by using the 488nm spectral line of an argon ion laser. A
constant number of up to 200 optical sections in Z-scan mode
was taken through each selected etched track at vertical
spacings of 0.642 micrometre. Successive reconstructions of Z-
sections were used to determine the following parameters: the
mean radius of the opening channel, the maximum diameter and
the length of the track, and the angle of the track wall to
the surface of the sample. The results show that tracks
produced by alpha particles differ from those induced by
protons. The radius of the opening channel of alpha-particle-
induced tracks ranges from 7.9 to 11 micrometre, whereas for
protons the same parameter ranges between 2.0 and 3.8
micrometre for a specific electrochemical etch procedure.



Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
85-89.

A simple method for overcoming some problems when observing
thick reflected biological samples with the confocal scanning
laser microscope

C. Rumio, M. Morini, J. R. Miani, I. Barajon & P. Castano
Institute of Human Anatomy, Via Mangiagalli 31, 20133 Milano,
Italy


SUMMARY

A simple device is described, which allows the range of depth
of scanning to be reduced when observing thick reflecting
biological specimens with a confocal scanning laser microscope
(CSLM). Thick histological sections of human skin and rat
brain stem were mounted between two coverslip ('sandwich
style') and the optical tomography was performed from both
sides by turning the 'sandwich' upside-down. The samples were
impregnated using standard Golgi-Cox, 'rapid Golgi' or other
silver methods. The ability to turn the sandwich upside-down
is particularly useful when the reflective structure inspected
is deep inside the section, i.e. near the lower surface of the
specimen, or when it is opaque to the laser beam of
excessively reflective.


Journal of Microscopy, Vol. 177, Pt. 1, January 1995, pp.
90-92.

Thick section preparation using a silicon-rubber-based sealant

H. Cox, C. Walker & C. V. Howard
Department of Orthopaedic & Accident Surgery, Royal Liverpool
University Hospital, Prescot Street, PO Box 147, Liverpool,
L69 3BX


SUMMARY

A method has been developed, using a silicon-rubber-based
sealant, which allows 2-3-mm-thick specimens to be maintained
in a protected fluid environment for a number of months,
without risk of dehydration. Following this, the specimen can
be retrieved, stained, embedded and sectioned further. For
example, 2-mm-thick sections of fixed unstained bone are
easily examined by means of epi-illuminated polarized light
and fluorescence microscopies using either conventional or
confocal optics. The method could easily be extended to other
tissues, for example brain tissue.



Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

A method to compensate for light attenuation with depth in
three-dimensional DNA image cytometry using a confocal
scanning laser microscope

A. Liljeborg, M. Czader & A. Porwit
Physics IV, The Royal Institute of Tech, S-100 44 Stockholm,
Sweden


SUMMARY

A method to compensate for attenuation of detected light with
increased depth of the collected optical section, and its
application in three-dimensional (3-D) DNA image cytometry is
described. The method is based on studying the stack of 2-D
histograms that ca be formed from each consecutive pair of
sections in a stack of optical serial sections. An attenuation
factor is calculated interactively and a new compensated
section series is computed. Formalin-fixed paraffin-embedded
rat tissue was stained with propidium iodide. Each cell
nucleus is extracted by thresholding and its total intensity
is calculated. The coefficient of variation (CV) of the total
intensity of all cells in each stack is computed. For
comparison the CV of the same cells is computed in the
uncompensated stacks. This study shows a significantly lower
CV for the compensated data, thus contributing to the accuracy
of DNA quantification in 3-D DNA image cytometry.


Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Microfractography of granitic rocks under confocal scanning
laser microscopy

M. Montoto, A. Martinez-Nistal, A. Rodriguez-Rey, N.
Fernandez-Merayo & P. Soriano
University of Oviedo, Department of Geology, Group of
Petrophysics, 33005 Oviedo, Spain


SUMMARY

Scanning laser microscopy, in the confocal mode (CSLM) has
been applied to a granitic rock to characterize its fissure
space. The technique provides a unique three-dimensional
picture of the rock microfractomography. CSLM is unique in
observing fine details of the fractographic network
(connectivity, tortuosity, etc.), its geometry and its
relation to other rock-forming components.
The fractographic images with standard fluorescence
microscopy are compared with those obtained with CSLM. The
examples presented emphasize the advantages of CSLM: three-
dimensional visualization of the microfractographic network,
crack connectivity, automatic evaluation of direction and
slope of fissures.
These studies are related to the migration of
radionuclides in the geosphere. The relations between
potentially water-conducting open fissures and the rock-
forming minerals provide a means of modelling the
'radionuclide retardation mechanism', a security factor in
their definitive storage in rock masses.


Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

PHOEBE, a prototype scanning laser-feedback microscope for
imaging biological cells in aqueous media

T. L. Wong, S. L. Sabato & A. Bearden
Division of Neurobiology, Department of Molecular & Cell
Biology, 229 Stanley Hall, University of California at
Berkeley, Berkeley CA 94720-3206, U.S.A.


SUMMARY

Based on the principle of laser-feedback interferometry (LFI),
a laser-feedback microscope (LFM) has been constructed,
capable of providing an axial (z) resolution of a target
surface topography of approximately 1nm and a lateral (x,y)
resolution of approximately 200nm when used with a high-
numerical-aperture oil-immersion microscope objective. LFI is
a form of interferometry in which a laser's intensity is
modulated by light re-entering the illuminating laser.
Interfering with the light circulating in the laser resonant
cavity, this back-reflected light gives information about an
object's position and reflectivity. Using a 1-mW He-Ne
(wavelength=632.8nm) laser, this microscope (PHOEBE) is
capable is obtaining 256x256-pixel images over fields from
10x10 micrometre to 120x120 micrometre in approximately 30s.
An electrochemical feedback circuit holds the optical
pathlength between the laser output mirror and a point on the
scanned object constant; this allows two types of images
(surface topography and surface reflectivity) to be obtained
simultaneously. For biological cells, imaging can be
accomplished using back-reflected light originating from small
refractive-index changes (} 0.02) at cell membrane/water
interfaces; alternatively, the optical pathlength through the
cell interior can be measured point-by-point by growing or
placing a cell suspension on a higher-reflecting substrate
(glass or silicon wafer). Advantages of the laser-feedback
microscope in comparison to other confocal optical microscopes
include: simplicity of the single-axis interferometric design;
the confocal property of the laser-feedback microscope (a
virtual pinhole), which is achieved by the requirement that
only light that re-enters the laser meeting the stringent
frequency, spatial (TEM00), and coherence requirements of the
laser cavity resonator mode modulate the laser frequency; and
the improved axial resolution, which is based on
interferometric measurement of optical amplitude and phase
rather than by use of a pinhole as in other types of confocal
microscopes.



Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Imaging periodic surface relief structures

J. T. Sheridan & T. O. Korner
TP 680, Institute for System Engineering and, Informatics
Science R&D, Joint Research Centre JRS/CCR, I-21020 Ispra
(VA), Italy


SUMMARY

Because of shadowing, multiple scatter and polarization
effects, the interpretation of images of grating with fine
periods, isolated deep structures, and multiple scattering
volume objects is seriously complicated. In this paper a
review of methods used to model such effects is presented.
Periodic surface relief gratings are of particular current
importance because of the possibility of producing calibration
samples using them. Several examples which illustrate
electromagnetic volume effects are examined. General trends
which help in validating the use of Fourier-transform-based
scalar transmittance theory are then indicated. The angular
spectrum approach, which can be used , together with a scatter
function generated using the rigorous electromagnetic theory,
to calculate coherent, partially coherent and confocal images
of volume objects, is also discussed.


Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Application of confocal laser microscopy and three-dimensional
Voronoi diagrams for volume and surface area estimates of
interphase chromosomes

R. Eils, E. Bertin, K. Saracoglu, B. Rinke, E. Schrock, Y.
Usson, M. Robert-Nicoud, E. H. K. Stelzer, J.-M. Chassery, T.
Cremer & C. Cremer
Institut fur Angewandte Physik, Albert-Uberle Strasse 3-5,
D-69120 Heidelberg, Germany


SUMMARY

This study demonstrates the use of Voronoi tessellation
procedures to obtain quantitative morphological data for
chromosome territories in the cell nucleus. As a model system,
chromosomes 7 and X were visualized in human female amniotic
fluid cell nuclei by chromosomal in situ suppression
hybridization with chromosome-specific composite probes. Light
optical serial sections of 18 nuclei were obtained with a
confocal scanning laser fluorescence microscope. A three-
dimensional (3-D) tessellation of the image volumes defined by
the stack of serial sections was then performed. For this
purpose a Voronoi diagram, which consists of convex polyhedra
structured in a graph environment, was built for each nucleus.
The chromosome territories were then described by three
morphological parameters, i.e. volume, surface area and a
roundness factor (shape factor). The complete evaluation of a
nucleus, including the calculation of the Voronoi diagram, 3-D
visualization of extracted territories using computer graphic
methods and parameterization was carried out on a Silicon
Graphics workstation and was generally completed within 5 min.
The geometric information obtained by this procedure revealed
that both X- and 7-chromosome territories were similar in
volume. Roundness factors indicated a pronounced variability
in interphase shape for both pairs of chromosomes. Surface
estimates showed a significant difference between the two X-
territories but not between chromosome 7-territories.


Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Forbidden light scanning near-field optical microscopy

H. Heinzelmann, B. Hecht, L. Novotny & D. W. Pohl
Institut fur Physik, Universitat Basel, Klingelbergstrasse 82,
4056 Basel, Switzerland


SUMMARY

Near-field optics (NFO) opens the door to light microscopy
techniques with resolutions well beyond the diffraction limit.
The richness of optical investigations is now applicable on a
near-molecular level. Among the novel scanning near-field
optical microscopy (SNOM) schemes, the most prominent
representatives are aperture SNOM and scanning tunnelling
optical microscopy (STOM or PSTM).
New experimental and theoretical work has to be performed
to study the phenomena specific to NFO. One such example is
the angular dependence of light emission in aperture SNOM. The
detection of radiation at angles greater than the critical
angle of total internal reflection alpha=arcsin(1/n), where n
is the sample refractive index, can represent a microscopy
scheme that combines the respective advantages of both
aperture SNOM and STOM. Recent experiments have demonstrated
the expected exponential dependence of light intensity on gap
width (for fixed emission angle). The decay length as a
function of alpha is in agreement with the Fresnel description
of the evanescent field when total reflection occurs at an
interface. These investigations were additionally motivated by
calculations based on the multiple multipole method.



Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Automated correction of linear deformation due to sectioning
in serial micrographs

T. Jansson, T. Gustavsson, M. Rydmark, C.-H. Berthold, R.
Pascher & T. Skoglund
Department of Applied Electronics, Chalmers University of
Technology, S-412 96 Goteborg, Sweden


SUMMARY

This paper describes an objective and automatic method for
detection and correction of sectioning deformations in
digitized micrographs, as well as an evaluation of the method
applied to light and electron microscopic images of semi-thin
and ultra-thin serial sections from brain cortex. The
detection is based on matching of image subregions and the
deformation model is bi-linear, i.e. two first-order
polynomials are used for modelling compression/expansion in
perpendicular directions. The procedure is applicable to
prealigned serial two-dimensional sections and is primarily
aimed at three-dimensional reconstruction of tissue samples
consisting of a large number of cells with random distribution
and morphology.


Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Comparison of three-dimensional imaging properties between
two-photon and single-photon confocal fluorescence microscopy

Min Gu & C. J. R. Sheppard
Physical Optics Department, School of Physics, The University
of Sydney, NSW 2006, Australia


SUMMARY

The imaging performance in single photon (1-p) and two-photon
(2-p) fluorescence microscopy is described. Both confocal and
conventional systems are compared in terms of the three-
dimensional (3-D) point spread function and the 3-D optical
transfer function. Images of fluorescent sharp edges and
layers are modelled, giving resolution in transverse and axial
directions. A comparison of the imaging properties is also
given for a 4Pi confocal system. Confocal 2-p 4Pi fluorescence
microscopy gives the best axial resolution in the sense that
its 3-D optical transfer function has the strongest response
along the axial direction.


Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Double pulse fluorescence lifetime imaging in confocal
microscopy

M. Muller, R. I. Ghauharali, K. Visscher, T. D. Visser & G. J.
Brakenhoff
Department of Molecular Cytology, University of Amsterdam,
Plantage Muidergracht 14, 1018 TV Amsterdam, The Netherlands


SUMMARY

A theoretical analysis of a new technique for fluorescence
lifetime measurement, relying on (near steady state)
excitation with short optical pulses, is presented.
Application of the technique to confocal microscopy enables
local determination of the fluorescence lifetime, which is a
parameter sensitive to the local environment of fluorescent
probe molecules in biological samples. The novel technique
provides good time resolution, since it relies on the laser
pulse duration, rather than on electronic gating techniques,
and permits, in combination with bilateral confocal microscopy
and the use of a (cooled) CCD, sensitive signal detection over
a large dynamic range. The principle of the technique is
discussed within a theoretical framework. The sensitivity of
the technique is analysed, taking into account:
photodegradation, the effect of the laser repetition rate and
the effect of non-steady-state excitation. The features of the
technique are compared to more conventional methods for
fluorescence lifetime imaging.


Journal of Microscopy, Vol. 177, Pt. 2, February 1995, pp. .

Near field imaging: some attempts to define an apparatus
function

D. Courjon
Laboratoire d'Optique PM Duffieux, Associe au CNRS URA-214,
UFR des Sciences et des Techniques, 25030 Besancon Cedex,
France


SUMMARY

Near-field microscopy is a promising new tool capable of
imaging details smaller than the wavelength. The mechanism of
imaging is analysed and an overview of the apparatus functions
which could be used to define an image quality criterion is
given.





From: rms-at-vax.ox.ac.uk
Date: Fri, 30 Jun 1995 11:14:29 +0100
Subject: Abstracts for J Microsc. Delete this message if not interested!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk
JTERLET-at-CEMMA.ADELAIDE.EDU.AU, microtoday-at-aol.com, ROSS.MACKENZIE-at-OX.AC.UK,
MICROSCOPY-at-AAEM.AMC.ANL.GOV
Message-ID: {00992A63.220570A6.10-at-vax.ox.ac.uk}





From: DRStadden:R_D:Armstrong
Date: 6-30-95 7:20am
Subject: FEG SEM for EDX?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: MICROSCO:CENTRAL
cc: DRStadden:R_D:Armstrong
Subj: FEG SEM for EDX?
------------------------------------------------------------------
A friend in academia is preparing to purchase an SEM to serve broad
interests in the sciences. He is wondering whether the field emission
gun instruments work as well for EDX of elements on the light side, such
as sodium and magnesium. Not too concerned with B,C,N,O and F. For
resolution and low KV work, he'd obviously like to have the FEG. Any
thoughts?

Dave Stadden

DRStadde+aCENTRAL%ARMSTRONG-at-MCIMAIL.COM
PH: 717-396-5109
FAX: 717-396-5865





From: rms-at-vax.ox.ac.uk
Date: Fri, 30 Jun 1995 13:19:47 +0100
Subject: More J. Microsc. Delete message now if unwanted!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk
JTERLET-at-CEMMA.ADELAIDE.EDU.AU, microtoday-at-aol.com, ROSS.MACKENZIE-at-OX.AC.UK,
MICROSCOPY-at-AAEM.AMC.ANL.GOV
Message-ID: {00992A74.A33AD506.13-at-vax.ox.ac.uk}


Journal of Microscopy
ABSTRACT FOR MAY 1995

Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 93-100.

Limits of electron probe formation

A. V. Crewe
The Enrico Fermi Institute and The Department of Physics, The University of Chicago,
5640 S. Ellis Avenue, Chicago, IL 60637, U.S.A.

Summary
The performance of many electron optical instruments is fundamentally limited by the
dimensions of the focused probe. This is true of the scanning electron microscope
and the scanning transmission electron microscope and, by inference, it may affect the
transmission electron microscope. There has been very little improvement over the
past few years and it seems reasonable to look for the explanation. It is possible to
arrive at some simple expressions for the limiting performances of conventional
instruments in a way that is independent of the design details and depends upon
practical limits of field strength. Experiment and theory also appear to be in
agreement with the fact that the limit for high-voltage instruments has been reached,
although there is still room for improvement for low voltages.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 101-106

Environmental scanning electron microscopy of marine aggregates

D. M. Lavoie,* B. J. Little,* R. I. Ray,* R. H. Bennett, M. W. Lambert, V. Asper &
R. J. Baerwald
*Code 7333, Code 7430, Naval Research Laboratory, Stennis Space Center, MS
39529, U.S.A., Center for Marine Science, University of Southern Mississippi, Stennis
Space Center, MS 39529, U.S.A., Department of Biology, University of New Orleans,
New Orleans, LA 70148, U.S.A.

Summary
Marine aggregates were examined for the first time in the hydrated state using an
environmental scanning electron microscope (ESEM). Sample preparation consisted
of fixation followed by rinsing with distilled water to remove excess salts and fixative.
Aggregates were continuously observed at resolutions comparable to conventional
scanning electron microscopy through stages of hydration, from completely immersed
to desiccated. Because no metallic coating is required, energy-dispersive X-ray
spectroscopy (EDXS) can be used to analyse rapidly constituent elements occurring
at low concentrations with no spectral interference. Subtle differences in mineral
particles were seen in both EDXS spectra and in direct observation of relative
hydration, reflecting apparent differences in mineralogy. ESEM enabled examination
of effects of desiccation and rehydration on individual particles composed primarily of
hydrated polymer and eliminated dehydration artefacts in delicate organisms.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 107-119

Mechanism of image formation for thick biological specimens: exit wavefront
reconstruction and electron energy-loss spectroscopic imaging

K. F. Han, J. W. Sedat & D. A. Agard
Graduate Group in Biophysics, Department of Biochemistry and Biophysics, and the
Howard Hughes Medical Institute, University of California at San Francisco, San
Francisco, CA 94143-0448, U.S.A.

Summary
With increasing frequency, cellular organelles and nuclear structures are being
investigated at high resolution using electron microscopic tomography of thick sections
(0.3-1.0 m). In order to reconstruct the structures in three dimensions accurately
from the observed image intensities, it is essential to understand the relationship
between the image intensity and the specimen mass density. The imaging of thick
specimens is complicated by the large fraction of multiple scattering which gives rise
to incoherent and partially coherent image components. Here we investigate the
mechanism of image formation for thick biological specimens at 200 and 300 keV in
order to resolve the coherent scattering component from the incoherent (multiple
scattering) components.
Two techniques were used: electron energy-loss spectroscopic imaging (ESI)
and exit wavefront reconstruction using a through-focus series. Although it is
commonly assumed that image formation of thick specimens is dominated by
amplitude (absorption) contrast, we have found that for conventionally stained
biological specimens phase contrast contributes significantly, and that at resolutions
better than 10 nm, superposed phase contrast dominates. It is shown that the
decrease in coherent scattering with specimen thickness is directly related to the
increase in multiple scattering. It is further shown that exit wavefront reconstruction
can exclude the microscope aberrations as well as the multiple scattering component
from the image formation. Since most of the inelastic scattering with these thick
specimens is actually multiple inelastic scattering, it is demonstrated that exit
wavefront reconstruction can act as a partial energy filter. By virtue of excluding the
multiple scattering, the 'restored' images display enhanced contrast and resolution.
These findings have direct implications for the three-dimensional reconstruction
of thick biological specimens where a simple direct relationship between image
intensity and mass density was assumed, and the aberrations were left uncorrected.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 120-124

Electrophoretic transfer of protein-pigment complexes from non-denaturing
gels to electron microscopy grids

S. Grber & M. K. Lyon
Molecular, Cellular and Developmental Biology, Campus Box 347, University of
Colorado, Boulder, CO 80309, U.S.A.

Summary

Different fractions of a mixture of protein-pigment complexes have been separated
from one another by non-denaturing gel electrophoresis. These complexes were
prepared for observation by inserting electron microscope grids directly into the
focused bands of the pigment-protein complexes and resuming electrophoresis for a
brief time, so that the complexes were deposited onto the grids. It was found that
complexes deposited from each band exhibited distinctly different appearances. It was
also found that the exact conditions of electrophoretic deposition onto the grids
affected the appearance of the complexes. The protein-pigment complexes were
characterized additionally by spectroscopy and denaturing gel electrophoresis.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 125-133

V(Z) curve formation of solid spherical microparticles in scanning acoustic
microscopy

K. I. Maslov,* P. V. Zinin, O. I. Lobkis* & T. Kundu
*Institute of Chemical Physics, Russian Academy of Sciences, Kosygin Str. 4, 117334
Moscow, Russia, Physics Department, Institute for Material Science and Structure
Research, University of Bremen, 28334 Bremen, Germany, Department of Civil
Engineering and Engineering Mechanics, University of Arizona, Tucson, AZ 85721,
U.S.A.

Summary
Information about the properties of materials in acoustic microscopy can be obtained
in the form of the V(Z) curves. The purpose of this paper is to present the theoretical
and experimental study of the V(Z) curve formation for solid spheres. It is shown that
an investigation of the position of different peaks in the V(Z) curves is useful to
determine the size and acoustical properties of a spherical particle.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 134-145

An ion microprobe study of the microchemistry of Ni-base superalloy-Al2O3
metal-matrix composites

K. K. Soni,* M. W. Tseng, D. B. Williams, J. M. Chabala,* Jianwei Li,* R. Levi-Setti*
& C. C. Bampton
*Enrico Fermi Institute and Department of Physics, The University of Chicago,
Chicago, IL 60637, U.S.A., Department of Materials Science and Engineering, Lehigh
University, Bethlehem, PA 18015, U.S.A., Rockwell International Science Center,
Thousand Oaks, CA 91358, U.S.A.

Summary
The Chemical microstructure of Ni-base superalloy/Al2O3 metal-matrix composites
(MMCs) has been studied by scanning ion microprobe microanalysis, using the
secondary ion mass spectrometry (SIMS) technique. The MMCs were fabricated using
the transient-liquid-phase bonding (TLP) process, with B-doped superalloy powder
as an interlayer. Boron was found to diffuse rapidly throughout the matrix to form
boride phases, mostly at the grain boundaries in the matrix. These borides contain
excess Cr (also Mo, Si, W) in comparison with the Ni alloy-matrix, but are depleted
in Ni (also in Al and Co). Carbides form at the grain boundaries as thin platelets and
inside the grains as fine particles. Chemical reaction occurs between the sapphire
fibre and the matrix; formation of NiAl2O4 spinel at the interface is suggested. This
interface reaction layer is friable and parts of it peel off during consolidation to become
inclusions in the matrix near the fibre/matrix interface.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 146-151

Estimation of individual feature surface area with the vertical spatial grid

L. M. Cruz-Orive* & C. V. Howard
*Stereology Unit, Department of Anatomy, University of Bern, Postfach 139, CH-3000
Bern 9, Switzerland, Department of Fetal and Infant Pathology, Royal Liverpool
Children's Hospital Alder Hey, Eaton Road, Liverpool L12 2AP, U.K.

Summary
The area of an individual bounded surface (e.g. the boundary of a properly sampled
cell) can be estimated from an isotropic uniform random stack of parallel sections, or
of non-invasive planar scans, using the well-known spatial grid. A standing problem
was to estimate the area of an individual bounded surface with an arbitrary degree of
accuracy from a vertical (i.e. not isotropic) stack of sections or scans. A new tool to
do this, called 'vertical spatial grid', is presented.



Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 152-159

Unbiased estimation of capillary length from vertical slices*

S. Batra, M. F Knig & L. M. Cruz-Orive
Department of Anatomy, University of Bern, P.O. Box 139, CH-3000 Bern 9,
Switzerland

Summary
Previous stereological approaches to estimate feature length include isotropic sections,
which tend to be inefficient for highly anisotropic structures such as skeletal muscle
capillaries, and semiparametric model-based methods, which require transverse and
longitudinal sections only, but are biased to a variable, unknown degree. The recent
method of vertical slices combines the advantages of both approaches, namely it is
unbiased, efficient and convenient. This study illustrates for the first time how to apply
the vertical slices method in biology by direct light microscopy and intersection
counting with a properly orientated cycloid test system. Neither image processing nor
confocal microscopy are used. The purpose of the study was to estimate capillary
length in the left ventricle of rat heart. Beyond this, a novel histochemical method
enables the staining of the venular capillary region in red and the arteriolar capillary
region in blue, and hence estimates their separate lengths. The vertical slices method
to estimate feature length seems to be a promising approach for biology.

* Paper read at the Sixth European Congress for Stereology, Prague, 7-10 September
1993.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 160-164

Computer-assisted method for simultaneous three-dimensional reconstruction
of highly magnified nerve endings and low-magnification contours of th spinal
cord

A. G. Liss
Department of Anatomy, Uppsala University and Department of Plastic and Hand
Surgery, Uppsala University Hospital, Uppsala, Sweden

Summary
Three-dimensional reconstructions of biological structures can be obtained by the use
of serial sections, tomography, confocal microscopy techniques and X-ray
crystallography. The earliest reconstructions were achieved manually, but semi-
automatic and automatic techniques are now available. Tissue studied by microscopy
contains structures of greatly varying dimensions. When producing a three-
dimensional reconstruction of the contours of the spinal cord and its white and grey
matter, greater magnification of the fine nerve endings may provide additional
information. However, reconstruction at different magnifications of the structures of
interest cannot be achieved automatically, but requires manual delineation of the
structures. A method is described in which a commercial program was used to
provide a wire-frame reconstruction which had been drawn by hand. The data were
processed further to obtain a realistic image which could be rotated to provide details
of the three-dimensional relationships of the spinal cord structures. This techique is
useful when relationships and details are otherwise difficult to comprehend due to
large size differences.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 165-181

Restoration of confocal images for quantitative image analysis

H. T. M. Van Der Voort* & K. C. Strasters
*Department of Molecular Cell Biology, University of Amsterdam, Plantage
Muidergracht 14, 1018 TV Amsterdam, The Netherlands, Pattern Recognition Group,
Faculty of Applied Physics, Delft University of Technology, Lorentzweg 1, 2628 CJ
Delft, The Netherlands

Summary
Quantitative studies of three-dimensional (3-D) structure of microscopic objects have
been made possible through the introduction of microscopic volume imaging
techniques, most notably the confocal fluorescence microscope (CFM). Although the
CFM is a true volume imager, its specific imaging properties give rise to distortions in
the images and hamper subsequent quantitative analysis. Therefore, it is a
prerequisite that confocal images are restored prior to analysis. The distortions can
be divided into several categories: attenuation of areas in the image due to self-
absorption, bleaching effects, geometrical effects and distortions due to diffraction
effects. Of these, absorption and diffraction effects are the most important. This
paper describes a method aimed at the correction of diffraction-induced distortions.
All the steps necessary in restoring confocal images are discussed, including a novel
method to measure instrumental properties on a routine basis. To test the restoration
procedure an image of a fluorescent planar object was restored. The results show a
considerable improvement in the z-resolution and no ringing artefacts. The relevance
of the method for image analysis is demonstrated by a comparison of results of
applying 3-D texture analysis to restored and unrestored images of a synthetic object.
Furthermore, the method can be successfully applied to noisy fluores cence images
of biological objects, such as interphase cell nuclei.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 182-184
Short Technical Note
A system for aligning video-recorded visual data and high-quality multichannel
recordings of simultaneously occurring signals

T. Lumsdon, M. Dickson & M. H. Gladden
Institute of Physiology, University of Glasgow, Glasgow G12 8QQ, Scotland, U.K.

Summary
An electronic circuit is described which provides a unique number for each video
frame, and which is superimposed on the frame as a 4-digit light-emitting diode
display using a second camera and videomixer. The number is also encoded as a
simple voltage waveform which is recorded separately together with electrical signals
giving information about events related to, and coincident with, the visual data, for
example nerve spike trains. Sophisticated computer analysis of signals can thus be
related accurately to microscopical visual information processed by video analysis.


Journal of Microscopy, Vol. 178, Pt. 2, May 1995, pp. 185-187
Short Technical Note
A low-cost digitized microscope stage for linear measurements

B. H. Rohitha
The Horticulture and Food Research Institute of New Zealand Ltd, Ruakura Research
Centre, Private Bag 3123, Hamilton, New Zealand

Summary
A low-cost device for making fine, but repetitive linear measurements under the
binocular microscope is described. Linear displacement is coupled on to digital
callipers which read the measurements in numerical values. The accuracy of the
measurements was such that their 95% confidence interval was within 0.01 mm when
read through the apparatus. The data were assimilated on a computer and then
transmitted into a PC for analysis.




From: rms-at-vax.ox.ac.uk
Date: Fri, 30 Jun 1995 13:21:15 +0100
Subject: Still more J. Microsc. abstracts. Delete this now if unwanted!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: rms-at-vax.ox.ac.uk
JTERLET-at-CEMMA.ADELAIDE.EDU.AU, microtoday-at-aol.com, ROSS.MACKENZIE-at-OX.AC.UK,
MICROSCOPY-at-AAEM.AMC.ANL.GOV
Message-ID: {00992A74.D76B6D86.18-at-vax.ox.ac.uk}

ABSTRACTS FOR MARCH AND APRIL 1995

Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 188-197.

Microscopy and formation of polymer/metal composites

D. Vesely, P. Disley & J. Pisacka Department of Materials Technology, Brunel University,
Uxbridge, Middlesex, UB8 3PH, U.K.

SUMMARY
A new class of materials, formed by dispersion of low-melting-point metal alloys in a polymer
matrix, has been studied from the point of view of microstructure, interfacial interaction and
mechanical properties. The phases in these composites were formed in the same way as for
polymer blends and were thus dependent on viscosity ratio, concentration, surface tension
and
interfacial interactions.
Metal alloys of tin and bismuth (Sn/Bi) were mixed with high-density polyethylene
(HDPE) and polystyrene (PS) at elevated temperatures. Some preliminary investigations of
lead and tin (Pb/Sn) alloys blended with HDPE, PS, polypropylene (PP), polyoxymethylene
(POM), polyethylene-terephtalate (PET), polymethylmethacrylate (PMMA) polycarbonate (PC),
polyvinylidenefluoride (PVDF) and polyvinylchloride (PVC) were also undertaken. The
composites were characterized by light and electron microscopy, image analysis, electrical
conductivity measurements and impact testing.
It is shown that the low-melting point metal alloys can be dispersed in polymers to a
submicrometre level by blending. The particle size distribution follows an exponential function,
which means that very fine as well as large particles are present. The equilibrium between
dispersion and coalescence is very rapidly established during mixing. The average particle
size can be controlled by the properties of the matrix, concentration of the metal and
processing conditions.
An investigation of interfaces revealed that in some cases a chemical interaction
between the metal and the polymer can occur. This is apparent by observation of
degradation, fluorescence and changes in mechanical properties.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 198-206

In-situ monitoring of fibre and matrix deformation in fibre-polymer composites

J.-P. Favre, M.-H. Auvray, P. Sigety, D. Lvque & C. Brianon Materials Department,
Office National d'Etudes et Recherches Arospatiales (ONERA), BP 72, 92322-Chtillon
Cedex, France

SUMMARY
Three methods, namely microphotoelasticimetry, Raman spectroscopy and surface microgrids,
are currently used at ONERA on fibre-reinforced polymers to measure the in-situ fibre or
matrix deformation. They provide the materials scientist with valuable indications on the early
damage growth. Recent results are given and information about the limitations of the
methods
are indicated.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 207-217

The effect of microstructure on flow promotion in resin transfer moulding reinforcement
fabrics

P. R. Griffin, S. M. Grove, F. J. Guild, P. Russell & J. Summerscales School of
Manufacturing, Materials and Mechanical Engineering, University of Plymouth, Plymouth,
Devon PL4 8AA, U.K., Department of Materials Science and Engineering, University of
Surrey, Guildford, Surrey GU2 5XH, U.K. and Department of Biological Sciences, University
of Plymouth, Plymouth, Devon PL4 8AA, U.K.

SUMMARY
The resin transfer moulding (RTM) process is becoming increasingly important for the
manufacture of continuous fibre-reinforced thermosetting resin matrix composites. The RTM
process is a closed mould technique which reduces volatile emissions relative to traditional
hand lay-up methods. The fibres, generally as several layers of fabric, are prepared as a
preform and laid in the closed mould. The resin is injected, at one or more points, and flows
through the mould to form the finished product. In the manufacture of high-performance
composite structures, the flow of resin is constrained by the high volume fraction of
reinforcement fibres required to achieve the performance. Commercial fabrics are becoming
available which are woven with specially designed mesoscale architecture to promote flow of
the resin. The flow rates in a series of such fabrics have been studied. The microstructures
of the resulting composites have been examined using brightfield optical microscopy. A
Quantimet image analyser was used to quantify the structures on both the mesoscale and the
microscale. The flow rate has been shown to be related to the presence of both large and
more modest sized pore space in the reinforcement architecture.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 218-229

Influence of matrix precursors on the microstructure and mechanical properties of C/C
composites

A. Figueiras, J. J. Fernndez, M. Granda, J. bermejo, E. Casal & R. Menndez Instituto
Nacional del Carbn, CSIC, Apartado 73, 33080 Oviedo, Spain

SUMMARY
The microstructure and properties of uni-directional pitch-based carbon-carbon composites
are explained in terms of the chemical composition of pitch precursors. Pitches are
characterized by standard procedures (elemental analysis, softening point and solubility tests),
extrography which is a simple and rapid silica gel absorption chromatographic technique,
Fourier transform infra-red and gas chromatography of the toluene-soluble fraction. Pitch
pyrolysis behaviour is monitored by hot-stage microscopy. The main microstructural features
of uni-directional composites from pitches and commercial PAN-based carbon fibres are
determined by light microscopy and scanning electron microscopy.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 230-241

Microstructure and thermo-mechanical stability of a low-oxygen Nicalon fibre

M. H. Berger, N. Hochet & A. R. Bunsell Centre des Materiaux Pierre Marie Fourt, Ecole
Nationale des Mines de Paris, BP 87, 91003 Evry cedex, France

SUMMARY
A new Nicalon SiC-based fibre, characterized by a low oxygen content (0.5% wt) has been
studied. The absence in this fibre of a continuous Si-C-O phase, which characterized the
previous NLM 202 series of fibres, induces larger mean sizes for the constituents: the fibre
is composed of -SiC grains 5-20 nm in diameter and turbostratic aggregates of carbon 2-5
nm in diameter. The fibre is seen to be stiffer at room temperature (E = 300 GPa) and
stronger due to a reduction in critical defects thanks to improvements in processing conditions.

The Young's modulus remains almost stable up to 1473 K in air and above this temperature
the core of the fibre exhibits continuous grain growth up to 1773 K, but without the
degradation that occurred in the previous generation of fibres. Fibre strength was seen to be
lowered when compared to room temperature values even when exposed in air to
temperatures of 1073 K. A comparable fall is not seen with the NLM 202 fibres until 1273 K
and this difference is attributed to the oxidation of the carbon-rich surface of the new fibre.
SiC is oxidized at higher temperatures, inducing, above 1473 K, the growth of a silica layer
on the surface, with defects at the glass/ceramic interface. The large discrepancies between
the good thermo-mechanical characteristics in inert atmosphere and the behaviour in air may
be reduced if a coating resistant to oxidation could be applied to the fibre.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 242 -250

The microstructure of experimental SiC fibre-reinforced yttrium magnesium
aluminosilicate (SiCf-YMAS) materials

J. Vicens, F. Doreau & J. L. Chermant LERMAT, URA CNRS 1317, ISMRA, 6 Boulevard
du Marchal Juin, 14050 Caen Cedex, France

SUMMARY
Two experimental SiC fibre-reinforced yttrium magnesium aluminosilicate (SiCf-YMAS)-type
ceramic-matrix composite (CMC) materials fabricated (i) by the glass process and (ii) by
chemical precursor infiltration have been studied by light microscopy, transmission electron
microscopy (TEM), high-resolution electron microscopy (HREM) and energy-dispersive X-ray
spectroscopy (EDS). The distribution of the fibres inside the composite as well as the
average
diameter of fibres have been determined by image analysis. The microstructure of the YMAS
matrices has been characterized by TEM observations. YMAS matrices are formed of two
main phases, cordierite and -yttrium silicate (Y2Si2O7). Two minor phases (mullite and
spinel) have been found to crystallize inside the cordierite and the yttrium silicate crystals.
Fibre-matrix interfaces have been observed in HREM. A thin turbostratic carbon layer (20-30
nm) has been imaged in both composites at the fibre-matrix interface. It crystallizes along
the matrix interface and grows inside the fibre, forming a diffuse interphase. The carbon layer
is believed to be the consequence of reaction between oxygen in the matrix and SiC
nanocrystals of the Nicalon fibres.



Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 251-272.

Environmental ageing effects in a silicon carbide fibre-reinforced glass-ceramic matrix
composite

K. P. Plucknett, S. Sutherland, A. M Daniel, R. L. Cain, G. West, D. M. R. Taplin & M.
H. Lewis School of Manufacturing, Materials and Mechanical Engineering, University of
Plymouth, Drake Circus, Plymouth PL4 8AA, U.K. and Centre for Advanced Materials,
University of Warwick, Coventry CV4 7AL, U.K.

SUMMARY
A silicon carbide fibre-reinforced glass-ceramic composite, based upon a BaO-MgO-Al2O3-
SiO2(BMAS) matrix, has been used for a study of microstructural stability (specifically
interface
stability) after environmental exposure at elevated temperature. Characterization of the as-
received material demonstrated the presence of a thin 'carbon-rich' interfacial layer between
fibre and matrix, as typically observed in glass-ceramic/silicon carbide fibre composite
systems. Samples have been subjected to heat-treatments in an oxidizing atmosphere at
temperatures between 723 and 1473 K, for up to 500 h. Intermediate-temperature ageing,
between 873 and 1073 K, results in strong fibre/matrix bonding, with consequent degradation
of strength and composite 'ductility'. This is due to oxidative removal of the carbon interfacial
layer and subsequent oxidation of the fibre surface, forming a silica bridge. Carbon is
retained
at higher ageing temperatures due to the formation of a protective surface oxide scale at
exposed fibre ends. Attempts to pretreat the BMAS composite at high temperature (1273 -
1473 K), designed to inhibit intermediate-temperature degradation via the formation of silica
plugs at exposed fibre ends, has given mixed results due to the high residual porosity content
in these materials, allowing paths of 'easy' oxygen ingress to be retained.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 272-278

Microstructure and evolution of a magnesium lithium aluminosilicate matrix composite

P. Ruterana, D. Kervadec, H. Maupas & J. L. Chermant LERMAT, URA CNRS 1317,
ISMRA, 6 Bd du Marchal Juin, 14050 Caen Cedex, France

SUMMARY
The microstructure of a magnesium lithium aluminosilicate glass ceramic composite has been
investigated by scanning electron microscopy and analytical transmission electron microscopy.

Attention was focused on the as-received material, showing that there is a non-uniform
distribution of the major silicate phases inside the matrix. The largest part is made of
spodumene type crystals containing more than 4 wt% Mg. A minor part of the matrix is made
of micrometre-sized crystallites of spodumene and cordierite. The spodumene is always
sensitive to the electron beam irradiation. The morphology of the amorphized spodumene
areas indicates that it may have crystallized during a later stage of the matrix formation, filling
the gaps between cordierite crystallites. The third component of the matrix is made of
carbon-rich areas. They can be as large as 10 m and they always include amorphous Mg-
rich silicates. However, they are mainly small (a few tens of nanometres in width) when
located at grain boundaries of spodumene crystals. In this case the turbostratic carbon
patches are also intimately mixed with an Mg-rich amorphous silicate. The interface between
the matrix and the fibres has also been analysed, its thickness changes from one to the other,
and it is sometimes empty due to decohesion. When it is filled, its outer part contains mainly
tubostratic carbon and the inner part is a mixture of silicon oxide and probably carbon. After
creep at 1373 K, the spodumene-type crystals are larger and they are no longer sensitive to
the electron beam. The cordierite areas appear to shrink and the amorphous patches which
were mixed with carbon transform into small crystallites (1-10 m). The areas next to the
fibres are found to extend irregularly into the matrix, probably as a result of a chemical
reaction.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 279-286

Synthesis of silicon carbide ceramics from a polysilastyrene at low temperatures

S. M. McMillan & R. J. Brook Department of Materials, University of Oxford, Oxford OX1
3PH, U.K.

SUMMARY
The conventional route for preparation of silicon carbide ceramics is by the use of
pressureless sintering, hot pressing, or hot isostatic pressing of silicon carbide starting
powders. High sintering temperatures (2073-2473 K) and the addition of sintering additives
are normally used to enhance densification. These sintering additives, however, form second
phases at grain boundaries which impair the mechanical properties of the material, particularly
at high temperatures. It is therefore desirable that new processing routes are developed that
overcome these difficulties. A proposed route is to use a polymeric pressure which can
provide a silicon carbide matrix as binding agent for silicon carbide powders, thus making the
requirement for high temperatures and sintering additives unnecessary. This paper reports
observations of the direct transformation of a polymeric precursor into amorphous Si-C, and
crystalline SiC at low temperatures, and the use of this precursor as a binder for the
production of SiC powder/precursor SiC composites.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 287-304

A comparison of the microstructure of silicon nitride-silicon carbide composites made
with and without deoxidized starting material

S. Turan & K. M. Knowles University of Cambridge, Department of Materials Science and
Metallurgy, Pembroke Street, Cambridge CB2 3QZ, U.K.

SUMMARY
Two different types of silicon carbide (SiC) matrix composites, with either 10 wt% or 20 wt%
silicon nitride (Si3N4) reinforcement, were fabricated to investigate the effect of pretreatment
on the resulting composite microstructure. The first type of composite was prepared from as-
received -SiC and -Si3N4 powders, while the second type was prepared from powder
compacts that had been deoxidized to eliminate surface silica on the powder particles. The
composites were hot isostatically pressed in tantalum cans at 2373 K for 1 h under a pressure
of 200 MPa. Density measurements showed that full theoretical density was achieved for the
composites prepared from the as-received powders, while much lower densities were
obtained for the composites prepared from the deoxidized green compacts. Almost all of the
-SiC transformed into -SiC, and almost all the -Si3N4 transformed into -Si3N4 in the
composites made from the as-received powders, while in the composites made from the
deoxidized material the -SiC remained untransformed and both -Si3N4 and -Si3N4
phases
were present in significant quantities. High-resolution transmission electron microscopy and
Fresnel fringe imaging were used to identify the grain boundary and interphase boundary
structure. Most interfaces were found to be covered with 1nm thick amorphous intergranular
films in the composites prepared from as-received powders, whereas most interfaces were
found to be free of such amorphous intergranular films in the composites prepared from the
deoxidized material. Taken together, the presence of intergranular films at the interfaces and
the results from density measurements are consistent with the densification and reverse
-SiC transformation taking place in the composites made from as-received powders by a
liquid-phase sintering route. An incomplete liquid-phase sintering mechanism is also able
to explain the microstructure observed in the composites made from the deoxidized material.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 305-312

Processing and properties of Al2O3/SiC nanocomposites

C. E. Borsa, S. Jiao, R. I. Todd & R. J. Brook Department of Materials, Oxford University,
Oxford, OX1 3PH, U.K.

SUMMARY
Alumina/SiC nanocomposites were produced by mechanical mixture of commercial powders.
The preparation steps involved the vigorous mixing of the powders and drying under
conditions where the homogeneous mixture was kept stable. Pressureless sintering of die-
pressed powders achieved reasonable densities ( 97% theoretical density) for 2.5wt% of SiC
on sintering at 2073 K. Higher SiC contents strongly reduced the sintered density. The use
of a more reactive alumina (finer matrix powder) gave similar results. Hot pressing at 1973
K/1 h/25 MPa produced high-density materials for SiC contents as high as 20 wt%.
Transmission and scanning electron microscopy analysis showed that the SiC particles were
well distributed and were situated both inside the grains and on the grain boundaries of the
alumina matrix. The SiC strongly inhibited grain growth in the matrix in keeping with the
Zener
model. The bend strength increased as the SiC content increased, a result partly explained
by the grain size refinement. The strength improvement of 20% over monolithic was
explained
in terms of the change to an intergranular fracture mode.



Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 313-319

Surface characteristics of Zirfon composite ultrafiltration membranes

S. Kuypers, I. Genn & R. Leysen VITO (Flemish Institute for Technological Research),
Materials Division, Boeretang 200, B-2400 Mol, Belgium

SUMMARY
This paper presents the first high-resolution field emission scanning electron micrographs of
the skin layers of a series of Zirfon ultrafiltration membranes. The flux through these
polymer-based composite membranes is known to be proportional to the inorganic-filler
content. Image analysis of skin surface micrographs revealed the skin surface pore
characteristics to be almost unchanged over a wide range of compositions. The inorganic
filler
was found to be present in the skin layer, where it modifies the polymer network and locally
reduces the skin thickness. This latter effect might be able to account for the observed flux
behaviour.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 320-330

Interfacial reactions in PZT/Pd/PZT sandwich structures

L. G. Yao & R. J. Brook Department of Materials, University of Oxford, Parks Road, Oxford
OX1 3PH, U.K.

SUMMARY
The interfacial reactions of palladium foil and lead zirconate-titanate (PZT) were studied using
samples with a sandwich structure in the temperature range 1373 - 1523 K and under
conditions where no lead is lost to the environment. The interfacial reactions were analysed
using scanning electron microscopy, energy-dispersive X-ray spectroscopy, X-ray diffraction
and wavelength-dispersive X-ray spectroscopy analysis. The density of the PZT powder
phase increased with increasing temperature and, when sintered above 1373 K, reached over
95% of the theoretical density of PZT. The weight loss of pellets was less than 0.8% when
sintered below 1523 K. The degree of interfacial reactions became more severe with
increasing temperature, as indicated by an expanding reacted region. The reaction at the
PZT
side of the Pd/PZT interface involved the decomposition of PZT into a monoclinic ZrO2 phase,
PbO and a lower -value (Pb(Zr Ti1- )O3 composition. Three distinguishable microstructures
exist on the Pd side when sintered below 1473 K: a thin layer of PbPd3 phase, a Pd-Pb solid
solution zone and an unreacted region. Only the cubic PbPd3 eutectic structure was found
when sintered above 1473 K. The oxidation of palladium occurred during interfacial reactions,
expedited by increasing temperature and resulting in the formation of the tetragonal PdO
phase and the hexagonal PbPdO2 phase. A model for the overall reaction is proposed
involving decomposition of the PZT, migration of PbO and diffusion of Pb into Pd foil.



Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 331-336

Microstructural investigation of colloidal silver embedded in glass

H. Hoffmeister & M. Dubiel Max Planck Institute of Microstructure Physics, Weinberg 2, D-
06120 Halle (Saale), Germany and Martin Luther University, Department of Physics,
Friedermann-Bach-Platz 6, D-06108 Halle (Saale), Germany

SUMMARY
Nanoparticulate composites consisting of very small silver particles embedded in near-surface
regions of glass were obtained by sodium-silver ion exchange. Colloidal silver is formed by
reduction of silver ions and aggregation of silver atoms in the glass matrix at elevated
temperatures. Owing to absorption bands in the visible region, the silver particles cause a
coloration of the glass that depends on their size and depth distribution. From high-resolution
electron microscopy imaging of lattice plane fringes of the silver particles, size-dependent
lattice contractions are deduced that are larger than those reported in the literature for
supported particles not interacting with a matrix. This effect is the more pronounced the
higher the annealing temperature is during the particle formation. The increased lattice
contraction is attributed to compressive stresses that arise during the ion exchange process
as well as during cooling after annealing.



Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 337-346

Spray forming of Al/SiC metal matrix composites

P. S. Grant, I. T. H. Chang & B. Cantor Oxford Centre for Advanced Materials and
Composites, Department of Materials, University of Oxford, Parks Road, Oxford OX1 3PH,
U.K.

SUMMARY
This paper describes the as-sprayed microstructure of a model Al-4wt%Cu/SiC particulate
(Al4Cu/Sicp) metal matrix composite (MMC) manufactured by spray forming, and the
relationship between microstructure and solidification conditions during manufacture.
Injection of SiCp into the melt atomization region during the spray forming of Al4Cu
results in significant SiCp incorporation into molten droplets during atomization, and relatively
little incorporation during flight to the substrate and at deposition. SiCp clustering is evident
in the Al4Cu droplets and results in clustering in the as-sprayed MMC deposit.
Matrix dislocation and precipitation microstructures are dependent upon local
solidification conditions during spray forming. Increased dislocation density and increased
quantity of fine-scale '-Al2Cu precipitation is found in the -Al(Cu) matrix where local
deposit
cooling rates are high, i.e. in the vicinity of the substrate/deposit interface and when increased
spray distances are used in manufacture. Lower dislocation density and increased quantity
of grain-boundary -Al2Cu is found where deposit cooling rates are relatively low, i.e. distant
from the substrate/deposit interface and at decreased spray distances. In all cases,
dislocation densities are higher in -Al(Cu)/SiCp interfacial regions than in the -Al(Cu)
matrix.
There is no evidence of -Al(Cu)/SiCp interfacial reaction in the as-sprayed condition
indicating that cooling rates during spray forming are sufficiently rapid to prevent reaction.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 347-356.

Microstructural origins of corrosion in a 20% SiCp/2124 aluminium alloy metal matrix
composite

D. Imeson & D. L. Bartlett Structural Materials Centre, DRA Farnborough, Farnborough,
Hants. GU14 6TD, U.K.

SUMMARY
The microstructure of a metal matrix composite (MMC) consisting of 20 wt% 3 m SiC
particles
in a 2124 Al alloy matrix has been examined and the relationship to corrosion pitting
processes investigated. Standard bulk corrosion tests show that the MMC forms a higher
density of pits than the unreinforced alloy, although the overall performance is similar as the
pits are shallower. In a new addition to conventional characterization techniques, transmission
electron microscopy samples have been directly subjected to 'flash' corrosion treatment and
subsequently examined. This techique is shown to be effective in studying the initiation of
pits. The SiC particles and the widespread intermetallic precipitates are shown to play little,
if any, role. A sparse population of features introduced during powder processing and
consolidation procedures, probably linked with strong magnesium segregation, is suggested
to be responsible for pit initiation.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 357-368

Microstructure and fracture behaviour of particle-reinforced metal-matrix composites

B. Derby Oxford Centre for Advanced Materials and Composites, Department of Materials,
Oxford University, Parks Road, Oxford OX1 3PH, U.K.

SUMMARY
The fracture behaviour of particle-reinforced metal-matrix composites is shown to be
controlled by a period of damage nucleation and evolution prior to final failure. The nucleation
of damage can be by reinforcement fracture or decohesion and the mode of damage is shown
to be controlled by the size of the reinforcement and segregation of alloying elements from
the matrix. The nucleation and growth of damage can be monitored by a number of
techniques. Acoustic emission and tomography are used here and the results are found to
be consistent with simple models of void growth.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 369-386

Spatially resolved electron energy-loss studies of metal-ceramic interfaces in transition
metal/alumina cermets

R. Brydson, H. Mllejans, J. Bruley, P. A. Trusty, X. Sun, J. A. Yeomans & M. Rhle
Department of Materials Science and Engineering, University of Surrey, Guildford GU2 5XH,
Max-Planck-Institut fr Metallforschung, Institut fr Werkstoffwissenschaft, Seestrasse 92,
70174 Stuttgart, Germany and department of Materials Science and Engineering, Lehigh
University, Bethleham, PA 18015, U.S.A.

SUMMARY
Composites consisting of an alumina matrix and 20 vol.% transition metal (Ni or Fe) particles,
prepared by hot pressing powder blends, have been studied using spatially resolved
transmission electron energy-loss spectroscopy (EELS), and, to a lesser extent, by high-
resolution electron microscopy (HREM). Particular attention was paid to the elucidation of the
chemical bonding mechanisms at the metal-ceramic interface; EELS spectra from interfacial
regions being obtained via a spatial difference technique. From both qualitative and
quantitative interpretation of EELS near-edge structures, as well as observed HREM images,
the data appear to be consistent with the presence of an Al-terminated alumina at the
interface and the formation of direct transition metal - aluminium bonds in Al(O3M) (M = Ni
or Fe) tetrahedral units, possibly as a result of the dissolution and interfacial reprecipitation
of Al during processing. These results correlate well with similar model studies on diffusion-
bonded Nb/Al2O3 interfaces and may, in the light of recent theoretical electronic structure
calculations, have implications for the resultant interfacial bond strength in such materials.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 387 - 398

Microstructure of a spray-formed Al/SiC composite

P. Vermaut & P. Ruterana LERMAT, URA CNRS No. 1317, ISMRA, 6 Boulevard du
Marchal Juin, 14050 Caen, Cedex, France

SUMMARY
The microstructure of an Al-6Cu-2Mn-0.45Mg-1(Ag,Ti,V,Zr,Cr) alloy, reinforced with 13 vol.%
SiC particles, made by spray deposition has been investigated by transmission electron
microscopy, high-resolution electron microscopy and electron diffraction X-ray spectroscopy.
Particular attention was focused on the influence of the reinforcement on the precipitation
sequence. Instead of the expected



precipitation sequence due to the high Cu/Mg ratio, there is an additional precipitate which
was previously observed in Al alloys containing silicon. This precipitate becomes predominant
at the T6 temper. The new precipitation sequence for this reinforced alloy is therefore




The precipitation of phase is believed to be due to the presence of SiC particles, and
seems to be correlated with the occurrence of large Mn-rich particles. Although expected,
no S phase precipitation is found to occur in the matrix grains. At the matrix grain boundaries,
small Al2Cu( ) and Al2CuMg (S), as well as Mn-rich precipitates are found. At the SiC
partical
surfaces, preferentially orientated Ag-rich and Mg-rich intermetallic precipitates are found.
They can coexist with amorphous patches containing oxygen enclosed in an irregularly
shaped
Al2Cu ( ) phase remaining from large crystalline areas which did not go into solution.




Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 399-406

X-ray microtomographic studies of metal matrix composites using laboratory X-ray
sources

P. M. Mummery, B. Derby, P. Anderson, G. R. Davis & J. C. Elliott Department of
Materials, University of Oxford, Parks Road, Oxford OX1 3PH, U.K. and Department of Child
Dental Health, London Hospital Medical College, Turner Street, London E1 2AD, U.K.

SUMMARY
X-ray microtomography (XMT) is a non-destructive technique that allows the internal
structure
of a material to be imaged by the spatial distribution of its linear X-ray absorption coefficients.

This paper demonstrates the use of XMT to investigate: (1) the distribution of TiB2
reinforcement in composites formed by powder processing; (2) the local void volume fraction
as a function of position in highly deformed regions of failed tensile specimens of SiC-
reinforced material allowing a valid damage parameter to be defined at high strains; (3)
absorption coefficients measured at different energies simultaneously using a multichannel
analyser which can sometimes be used to separate linear absorption changes due to (a)
density variations and (b) compositional variations in individual voxels; and (4) the use of
sequential sections to provide a three-dimensional representation of the failed specimens.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 407-413

The use of a single fibre test technique to investigate interfacial phenomena in SiC/Ti3
Al-based composites

F. Brisset & A. Vassel Direction des Matriaux, ONERA, BP 72, 92322 Chtillon Cedex,
France

SUMMARY
A study of the chemical compatibility of a Ti3Al-based alloy (Ti-24A1-10Nb, at.%) with two
silicon carbide continuous fibres (SCS-6, SM 1240) has been conducted. Owing to the
difficulty in processing intermetallic matrix composites, this type of material has been
simulated in the present work by sputtering a thin titanium aluminide layer onto fibres and heat
treating at temperatures representative of fabrication conditions. The degradation of the fibre
strength due to its interaction with the matrix was correlated with analytical studies of the
fibre/matrix interface using a combination of SEM, TEM, EELS and a submicrometre ion
probe.


Journal of Microscopy, Vol. 177, Pt. 3, March 1995, pp. 414-423

Ion microprobe studies of reactions in squeeze-cast aluminium alloy matrix composites

K. K. Soni, H. G. Kang, P. S. Grant, B. Cantor, A. G. Adriaens, K. L. Gavrilov, R.
Mogilevsky, R. Levi-Setti, M. W. Tseng & D. B. Williams Enrico Fermi Institute and
Department of Physics, The University of Chicago, Chicago, IL 60637, U.S.A., Oxford Centre
for Advanced Materials and Composites, Department of Materials, University of Oxford, Oxford

OX1 3PH, U.K. and Department of Materials Science and Engineering, Lehigh University,
Bethlehem, PA 18015, U.S.A.

SUMMARY
An ion microprobe with high lateral resolution has been used to study the chemical reactions
at the fibre/matrix interface of metal-matrix composites. During the squeeze-casting
process,
the Al-Si-Mg matrix reacts with the preform made of Saffil fibres (96% Al2O3, 4% SiO2). The
reaction occurs mainly between the silica binder and Mg from the matrix according to SiO2 +
2Mg = 2MgO + Si. A continuous layer of MgO was formed around the fibres, even on
surfaces that were not covered by the silica binder. Possible reasons are discussed for the
formation of MgO in areas where binder coating was missing. In such areas, Mg reduces
SiO2 that is contained in the fibre. However, the fibres (Al2O3) are not attacked by Mg. In
the
isolated case of fibres that were completely uncoated, no reaction products were observed at
the interface. The presence of silica binder seems to be an essential requirement for this
reaction to occur. When squeeze-casting is performed with sufficiently high melt temperature,
Al from the matrix also reduces silica.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 1-6

Identification and surface structure of crystalline cellulose studied by atomic force
microscopy

L. Kuutti, J. Peltonen, J. Pere & O. Teleman VTT, Biotechnology and Food Research, PO
Box 1503, FIN-02044 Espoo, Finland and Department of Physical Chemistry, bo Akademi
University, Porthaninkatu 3-5, FIN-20500 Turku, Finland

SUMMARY
A combination of molecular modelling and atomic force microscopy (AFM) techniques was
used to study the surface structure of crystalline cellulose. Two-dimensional Fourier analysis
of the AFM raw data gave crystal parameters as well as a highly filtered inverse-transformed
image. Molecular modelling was used to generate Connolly surfaces based on electron
diffraction data for crystalline cellulose. The modelled surfaces were used to interpret the
experimental AFM images. Monoclinic (110) crystal faces were identified. The method used
enables the structural analysis of cellulose surfaces at the molecular level, where all biological
processes involving cellulose take place.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 7-13

Atomic force microscopy under liquid: a comparative study of three different AC mode
operations

T. M. H. Wong & P. Descouts Group of Applied Physics, University of Geneva, CH-1211
Geneva 4, Switzerland

SUMMARY
The image contrast mechanism of three newly proposed AC mode operations under liquid in
three different frequency ranges is presented. They all rely on a strong repulsive force to
damp the cantilever and the tip still 'touches' the sample surface. A direct comparison of the
three different modes of operation with the conventional DC mode technique using the same
gold sample under isopropanol was conducted. It was found that all the three AC modes
exerted a much smaller lateral force than the DC mode although the normal loads were of the
same order of magnitude. The suitability of such techniques in imaging physisorbed systems
on hard substrates (such as soft biological samples) and the prospect of a true non-contact
repulsive mode operation are discussed.


Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 14-19

A scanning near-field optical microscope having scanning electron tunnelling
microscope capability using a single metallic probe tip

Y. Inouye & S. Kawata Department of Applied Physics, Osaka University, Suita, Osaka 565,
Japan

SUMMARY
A new microscope system that has the combined capabilities of a scanning near-field optical
microscope (SNOM) and a scanning tunnelling microscope (STM) is described. This is
achieved with the use of a single metallic probe tip. The distance between the probe tip and
the sample surface is regulated by keeping the tunnelling current constant. In this mode of
operation, information about the optical properties of the sample, such as its refractive index
distribution and absorption characteristics, can be disassociated from the information
describing its surface structure. Details of the surface structure can be studied at resolutions
smaller than the illumination wavelength. The performance of the microscope is evaluated by
analysing a grating sample that was made by coating a glass substrate with gold. The results
are then compared with the corresponding SNOM and STM images of the grating.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 20-27

Three-dimensionally resolved NAD(P)H cellular metabolic redox imaging of the in situ
cornea with two-photon excitation laser scanning microscopy

D. W. Piston, B. R. Masters & W. W. Webb School of Applied and Engineering Physics,
Cornell University, Clark Hall, Ithaca, NY 14853, U.S.A., Department of Anatomy and Cell
Biology, Uniformed Services University of the Health Sciences, 4301 Jones Bridge Road,
Bethesda, MD 20814, U.S.A. and Department of Molecular Physiology and Biophysics,
Vanderbilt University Medical Center, 702 Light Hall, Nashville, TN37232, U.S.A.

SUMMARY
Three-dimensional maps of cellular metabolic oxidation/reduction states of rabbit cornea in
situ were obtained by imaging the fluorescence of the naturally occurring reduced pyridine
nucleotides (both reduced nicotinamide-adenine dinucleotide, NADH, and reduced
nicotinamide-adenine dinucleotide phosphate, NADPH, denoted here as NAD(P)H.
Autofluorescence images with submicrometre lateral resolution were obtained throughout the
entire 400 m thickness of the cornea. Two-photon excitation scanning laser microscopy with
near-infrared excitation provided high fluorescence collection efficiency, reduced
photodamage, and eliminated ultraviolet chromatic aberration, all of which have previously
degraded the visualization of pyridine nucleotide fluorescence. Sharp autofluorescence
images of the basal epithelium (40 m within the cornea) show substantial subcellular detail,
providing the ability to monitor autofluorescence intensity changes over time, which reflect
changes in oxidative metabolism and cellular dynamics necessary for maintenance of the
ocular surface. The autofluorescence was confirmed to be mostly of NAD(P)H origin by
cyanide exposure, which increased the fluorescence from all cell types in the cornea by about
a factor of two. Autofluorescence images of individual keratocytes in the stroma were
observed only after cyanide treatment, while in the predominant extracellular collagen (} 90%
of the stromal volume), fluorescence was not distinguished from the background. Observation
of keratocyte metabolism demonstrates the sensitivity made available by two-photon
microscopy for future redox fluorescence imaging of cellular metabolic states.


Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 28-36

A hybrid scanning force and light microscope for surface imaging and three-
dimensional optical sectioning in differential interference contrast

A. Stemmer Marine Biological Laboratory, Woods Hole, MA 02543, U.S.A.

SUMMARY
The design of a scanned-cantilever-type force microscope is presented which is fully
integrated into an inverted high-resolution video-enhanced light microscope. This set-up
allows us to acquire thin optical sections in differential interference contrast (DIC) or
polarization while the force microscope is in place. Such a hybrid microscope provides a
unique platform to study how cell surface properties determine, or are affected by, the three-
dimensional dynamic organization inside the living cell.
The hybrid microscope presented in this paper has proven reliable and versatile for
biological applications. It is the only instrument that can image a specimen by force
microscopy and high-power DIC without having either to translate the specimen or to remove
the force microscope. Adaptation of the design features could greatly enhance the suitability
of other force microscopes for biological work.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 37-41

Preliminary confocal scanning laser microscopy study of fluid inclusions in quartz

N. Petford, J. A. Miller & A. H. Rankin School of Geological Sciences, Kingston
University, Kingston KT1 2EE, U.K. School of Geological Sciences, Kingston University,
Kingston KT1 2EE, U.K. and Bullard Laboratories, Department of Earth Sciences, University
of cambridge, Madingley Road, Cambridge CB3 0EZ, U.K.

SUMMARY
The initial results of the first dedicated confocal scanning laser microscopy (CSLM) study of
fluid inclusions in quartz are presented. CSLM imaging of a large inclusion shows the quartz
crystal to contain numerous small ( {1 m), highly reflective inclusions arranged along planes
in at least two directions that are not readily visible in transmitted light. The technique allows
measurements to be made of the angular intersection and orientation of the planes in both
two
and three dimensions. Results suggest that larger inclusions (} 10 m) occur where two
planes of small inclusions intersect, and that the shape of the large inclusions is controlled by
the angular relationship between intersecting planes.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 42-47

STM imaging of metal-coated cell plugs of the archaeobacterium Methanospirillum
hungatei GP1

W. Xu, B. L. Blackford, P. J. Mulhern, M. H. Jericho, M. Firtel & T. J. Beveridge
Department of Physics, Dalhousie University, Halifax, Nova Scotia B3H 3J5, Canada and
Department of Microbiology, College of Biological Science, University of Guelph, Guelph,
Ontario, N1G 2W1, Canada

SUMMARY
Scanning tunnelling microscopy (STM) images of Pt/Ir- and Pt/Ir/C-coated cell plugs of
Methanospirillum hungatei showed paracrystalline structures with P6 symmetry and an 18-nm
lattice constant, in agreement with electron microscopy studies. The three-dimensional STM
images unambiguously distinguished the two morphologically different proteinaceous plug
assemblies and led to an improved understanding of the natural internal organization of whole
plugs. Tip convolution effects and the grain size of the metal coating complicated
interpretation of finer structures. We discuss possible imaging mechanisms to explain
observations in which part of the film was removed but the remaining part of the structure was
still imaged reproducibly.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 48-55

Maximum entropy reconstruction of compositional depth profiles from electron probe
microanalysis data

G. C. Smith, D. Park & O. Cochonneau Shell Research Ltd, Thornton Research Centre, PO
Box 1, Chester CH1 3SH, U.K. and Departement de Genie Mathematique, INSA de Rouen,
BP 08, 76131 Mont Saint Aignan, France

SUMMARY
Electron probe microanalysis (EPMA) is a powerful method for the quantitative determination
of the elemental composition of micro-regions of a sample surface. Here, we report on the
development of a method of reconstructing compositional depth profiles in thin films from
EPMA data measured over a range of electron beam energies, using maximum entropy data
processing. The method gives quantitative information on film compositions up to
approximately 1 m in depth, with a lateral spatial resolution of approximately 1 m. The
method is tested using both simulated data and measured experimental data from well-
characterized model sample structures.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 56-65

Monitoring cutting forces with an instrumented histological microtome

A. Willis & J. F. V. Vincent Centre for Biomimetics, The University, Reading, U.K.

SUMMARY
A modification to the knife mounting of a histological microtome is described which can sense
the load on the knife when a section is being cut. The performance of this microtome is
described in general terms. The variation in force on the knife can be used to give
information
about the texture of the sample being cut.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 66-85

The biological reality of the interlacunar network in the embryonic, cartilaginous,
skeleton: a thiazine dye/absolute ethanol/LR White resin protocol for visualizing the
network with minimal tissue shrinkage

D. M. Lawton, W. B. Oswald & J. McClure Department of Pathological Sciences, University
of Manchester, Stopford Building, Oxford Road, Manchester M13 9PT, U.K.

SUMMARY
Third toe phalanges of chicks aged 8-13 days in ovo and 7-day post-natal rat femoral growth
plate were examined to determine whether the interlacunar network (IN), a structure with no
lipoprotein membrane component or cytoplasmic organelles, is a genuine component of young
growth cartilage. In chick phalanges dehydrated by 70% (v/v) ethanol and LR White resin,
variable metachromatic staining of the interlacunar network by toluidine blue and red staining
by picro-Sirius red indicate the presence of glycosaminoglycans and collagen. The network
in phalanges dehydrated by 80% (v/v) ethanol apears little different; however, the network is
much less widely detectable in phalanges dehydrated by 90% (v/v) ethanol and, after
dehydration by absolute ethanol, is almost completely undetectable. In contrast, when the
young cartilage is permeated by a thiazine dye such as toluidine blue, using a solution of dye
in the aldehyde fixative, the network is widely detectable, following dehydration by absolute
ethanol, both in chick phalanges and in rat growth plate. Comparison of projected areas
shows that the extent to which whole chick feet are found to have shrunk, by the time that
they are photographed under LR White resin, is determined principally by the extent of
dehydration, by 70% (v/v) or absolute ethanol; post-shrinkage areas are 33% or 35% of areas
measured in buffer for 70% (v/v) ethanol/LR White resin and 71% or 75% for absolute
ethanol/LR White resin (the higher value in each is for the toluidine blue treatment). The
network is thus present in radically shrunk tissue, but, significantly, is also fully represented
in tissue shrunk by only a conventional margin and is therefore not produced as an artefact
by exceptional tissue shrinkage as has been suggested.



Journal of Microscopy, Vol. 178, Pt. 1, April 1995, pp. 86-92

Inherent bias in correlation averaged images

J. T. Woodward, C. Kono, L. L. Madsen & J. A. Zasadzinski Department of Physics,
Materials Research Laboratory, and Department of Chemical and Nuclear Engineering,
University of California, Santa Barbara, CA 93106, U.S.A.

SUMMARY
The correlation averaging algorithm frequently used to enhance micrographs of repeating
structures contains an inherent bias that favours images with larger pixel values or positive
noise levels. This bias not only skews the composite image toward higher pixel values, but
also distorts the image by increasing the value of high-valued pixels more than that of low-
valued pixels. These errors are especially important in scanning probe microscopy images
where the pixel value reflects a distinct height. A similar algorithm that uses a structure
function in place of the correlation function eliminates this bias.





From: mohan-at-mtu.edu (Mohan Krishnamurthy)
Date: Fri, 30 Jun 1995 09:12:51 -0500
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199506301301.JAA03452-at-mmserver.mm.mtu.edu}
X-Sender: mohan-at-mmserver.mm.mtu.edu
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

unsubscribe
mohan-at-mtu.edu
krishnam-at-csss.la.asu.edu






From: Carlo Montemagno :      cmontema-at-hawk.ce.nd.edu
Date: Fri, 30 Jun 1995 13:16:19 -0500
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: Krzysztof Hubner :      zahubner-at-cyf-kr.edu.pl
Date: Fri, 30 Jun 1995 17:27:08 +0200 (METDST)
Subject: short glass fibre composite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1407615833.63628-at-qmgate.anl.gov}
"Stadden, David R" {DRSTADDE+aCENTRAL%Armstrong-at-mcimail.com}
X-Mailer: Mail*Link SMTP-QM 3.0.2

RE} FEG SEM for EDX? 6/30/95

Go for the FEG! The working distance is usually closer so you can do nice
imaging as well as EDS. The only limitation is the detector window material
and the Kv for the beam.
Obviously if you want to see sodium you need at least 1 kv, maybe 2kv to
excite the X-rays.

--------------------------------------
#012#
To: MICROSCO:CENTRAL
cc: DRStadden:R_D:Armstrong
Subj: FEG SEM for EDX?
------------------------------------------------------------------
A friend in academia is preparing to purchase an SEM to serve broad
interests in the sciences. He is wondering whether the field emission
gun instruments work as well for EDX of elements on the light side, such
as sodium and magnesium. Not too concerned with B,C,N,O and F. For
resolution and low KV work, he'd obviously like to have the FEG. Any
thoughts?

Dave Stadden

DRStadde+aCENTRAL%ARMSTRONG-at-MCIMAIL.COM
PH: 717-396-5109
FAX: 717-396-5865


------------------ RFC822 Header Follows ------------------
Received: by qmgate.anl.gov with SMTP;30 Jun 1995 07:40:56 -0600


Dear Friends

I have the following problem.
I was given some samples of a composite: short glass fibre -
transparent resin used as matrix.
I am supposed to examine under microscope the orientation of
these fibres.

The difference in grey level between the fibre and the resin is
very small; in practice only grey boundaries of the fibre - resin
system are visible.

The attempt to use various optical techniques (polarised light,
phase contrast etc.) to intensify the contrast between fibre and resin gave
a negative result. Etching in CrO2, HF was no good either.

And hence my question - maybe you now some
other means of preparation of the samples and of intensifying the
contrast between fibre and resin:
- chemical etching,
- image processing by means of mathematical morphology.

Thank you your kind reply.
Accept my best regards.


Krzysztof Jan Hubner {zahubner-at-cyf-kr.edu.pl}
Foundry Research Institute
Research Department - Structural and Physical Research Laboratory
Zakopianska 73 Call +48 12 605022 ext. 356
30-148 KRAKOW - POLAND Fax +48 12 665478 :-)






From: Richard Lee :      richard_lee-at-qmgate.anl.gov
Date: 30 Jun 1995 11:14:30 -0600
Subject: Re: FEG SEM for EDX?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1407611436.28988-at-qmgate.anl.gov}
"Stadden, David R" {DRSTADDE+aCENTRAL%Armstrong-at-mcimail.com}
X-Mailer: Mail*Link SMTP-QM 3.0.2




From: Richard Lee :      richard_lee-at-qmgate.anl.gov
Date: 30 Jun 1995 10:00:27 -0600
Subject: Re: FEG SEM for EDX?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

RE} FEG SEM for EDX? 6/30/95

Go for the FEG! The working distance is usually closer so you can do nice
imaging as well as EDS. The only limitation is the detector window material
and the Kv for the beam.
Obviously if you want to see sodium you need at least 1 kv, maybe 2kv to
excite the X-rays.

--------------------------------------
#012#
To: MICROSCO:CENTRAL
cc: DRStadden:R_D:Armstrong
Subj: FEG SEM for EDX?
------------------------------------------------------------------
A friend in academia is preparing to purchase an SEM to serve broad
interests in the sciences. He is wondering whether the field emission
gun instruments work as well for EDX of elements on the light side, such
as sodium and magnesium. Not too concerned with B,C,N,O and F. For
resolution and low KV work, he'd obviously like to have the FEG. Any
thoughts?

Dave Stadden

DRStadde+aCENTRAL%ARMSTRONG-at-MCIMAIL.COM
PH: 717-396-5109
FAX: 717-396-5865


------------------ RFC822 Header Follows ------------------
Received: by qmgate.anl.gov with SMTP;30 Jun 1995 07:40:56 -0600
"Stadden, David R" {DRSTADDE+aCENTRAL%Armstrong-at-mcimail.com}
X-Mailer: Mail*Link SMTP-QM 3.0.2







From: Richard Crang :      rcrang-at-pop.life.uiuc.edu
Date: Thu, 29 Jun 1995 12:51:00 -0500
Subject: Containing Unicells for Freeze-Substitution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X400-Received: by /PRMD=ABTLAB/ADMD=BT/C=US/; Relayed; Thu,
29 Jun 1995 12:51:00 -0500
Alternate-Recipient: prohibited
{MICROSCOPY-at-AAEM.AMC.ANL.GOV} ,
Admin {/O=PRDAP/OU=PRDAPP01/S=Admin/-at-x400.pprd.abbott.com}
Message-Id:
{"00005BAC.MAI*/DD.RFC-822=rcrang(a)pop.life.uiuc.edu/O=PRDSMTP/PRMD=ABTLAB/ADMD=BT/C=US/"-at-MHS}
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; CHARSET=US-ASCII
Content-Transfer-Encoding: 7BIT
Ua-Content-Id: CSI NC V3.0
X400-Content-Type: P2-1984 (2)
X400-Mts-Identifier: [/PRMD=ABTLAB/ADMD=BT/C=US/;950629175100]

microscopy studies and would like to have some of them prepared by means of
freeze substitution. They are in an osmotically adjusted buffer medium as a
cell suspension and, while we can centrifuge them (gently) to condense them
and even freeze them in chilled propane, they will disperse during the
substitution process, and particularly upon warming. Thus, they will be
lost upon the exchange of fluids. We are using a Reichert CS-Auto FS
system. Any suggestions how the material (or any unicellular preparations
for that matter) can be handled to contain them during freeze substitution?
Maintaining the cells on agar is not possible due to the need to keep them
in an osmotic medium prior to freezing. Suggestions greatly appreciated.



**************************
Richard F. E. Crang
Professor of Plant Biology
University of Illinois
(217) 244-3143
**************************





From: wise-at-vaxa.cis.uwosh.edu
Date: Fri, 30 Jun 1995 10:40:25 +0000
Subject: Yet more legal/ethical drivel

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There are several points I would like to add to the discussion
about unfairness of tax-supported (and tax-exempt) universities competing
against the heavily-taxed industrial sector.

--The tax breaks (or subsidies) given to the industrial sector since the
beginning of the Regan era have substantially shifted the tax burden from
businesses onto private citizens (mostly the middle class--where virtually
all university professors reside). The American business community has
never seen a more favorable tax or regulatory environment in the history of
our nation. The present Congress is charging ahead to further reduce taxes
and safety, health, environmental and fair business practice regulations as
they pertain to the industrial sector, thus further improving the business
environment. I submit that American business is not being taxed and
regulated to death.

--Public funding for universities has declined steadily for the last 20
years. Academic departments typically are only able to replace 1 (maybe
1.5) out of every three retirements. Faculty numbers are decreasing while
student enrollments stay the same or increase. A "four-year degree" takes
5 years at most schools because students, in spite of paying higher tuition
every year, cannot get into the classes they need for graduation. As an
example, look at how much public support the California public higher
education system enjoys from its citizenry. The University of Wisconsin
System will receive a 5.1% ($43 million) reduction in state moneys in the
next fiscal year. Pay raises for UW faculty (who are forbidden by state
law from unionizing) have been at or below inflation for most of the last
two decades. The notion of bottomless public coffers being used to drive
private industry out of business runs at odds with the current state of
financial support of public higher education in this country.

--My understanding is that most of the private EM testing labs that have
closed or down-sized recently are the result of a precipitous drop-off in
the asbestos testing business--not unfair competition from tax-supported
public universities.

--Dr. Garber's concern about the liability issue seems somewhat misplaced.
American universities are responsible for the health and safety of
literally millions of students and employees every day, any one of which
could sue the university at any time, for any reason. To state that "this
activity (product testing) puts the entire university at risk in ways that
are never even considered" and that "the real losers are the taxpayers, who
now have to carry the burden or risk of the folly of the commercial
activities being conducted out of the university's laboratories" implies
that this huge industry (i.e. public higher education), that predates the
founding of our country, has never faced a law suit. University
administrators are fully aware of the litigious society in which we live.
Hell, we train the lawyers!

--The UW Oshkosh Hitachi 2460N SEM with Noran Voyager EDS was paid for (in
total) by a gift from an anonymous, local industrialist. Does that exempt
us from NSF Notice 91 and those who interpret it to cover all
instrumentation housed at a public university (whether NSF-supported or
not)? What if our private benefactor asked for some free beam time (on the
machine he/she paid for)? Would I have to turn him/her away?

--In our situation, it was members of the private sector who came to us and
asked if they could rent our SEM (not the other way around). When we told
them "Yes, but at the same rate as you would have to pay a private lab"
they bolted, pulled an old SEM out of mothballs, and I haven't heard from
them since. They wanted access to a state-of-the-art SEM/EDS without
having to pay the going rate. Without the cooperation, and even
encouragement, of members of the private sector, universities would have no
market for the testing services which some would say unfairly compete
against other members of the private sector.

Let me close by saying that we provide no services at UW Oshkosh
that compete with the private sector. We did look into the possibility
once when specifically approached by industry, but have never pursued it of
our own volition. However, I do take a little umbrage at the notion that
fat university professors are using tax-payer's dollars to sweeten their
own nests while the tax-paying private sector carries the fiscal burden for
it all. It's a competitive world out there, guys. If you don't believe
me, apply for a tenure-track position somewhere.

Bob

Robert R. Wise, PhD
Director, UWO Electron Microscope Facility
Department of Biology
University of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404
(414) 424-1101 fax
wise-at-vaxa.cis.uwosh.edu






From: Christine.A.Snay-at-Dartmouth.EDU (Christine A. Snay)
Date: 30 Jun 95 15:05:00 EDT
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {6301701-at-schlitzen.Dartmouth.EDU}

Subscribe
Christine A. Snay-at-dartmouth.edu




From: AWBlackwoo-at-aol.com
Date: Fri, 30 Jun 1995 16:13:18 -0400
Subject: Glass Fiber Composites

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dr. Krzysztof Hubner has raised some questions about examination of
transparent resin matrices reinforced with short glass fibers.

I must say that most of our experience has been with systems which produce
very good contrast under polarized light. They are not black/white
differences, however, they are shades of gray.

There is another approach to this problem which has produced very good
results, also. The use of an oxygen plasma will rapidy remove the matrix
while leaving the glass fibers untouched. Scanning electron microscopy can
then be used to image the fibers and their orientation, with as much contrast
as one might desire.

For this use, the nondirectional plasma of the barrel type of etcher would be
better than the directional plasma of the parallel plate type.

Yes, my employer, Structure Probe, sells both kinds of etchers through its
SPI Supplies Division, and further information on these units can be obtained
from SpiSupp-at-aol.com or by contacting me.

Andy Blackwood

Andrew W. Blackwood, Ph.D.
Structure Probe, Inc.
63 Unquowa Road
Fairfield, CT 06430-5015
Ph: 1 203 254 0000
FAX: 1 203 254 2262
e-mail: AWBlackwoo-at-aol.com





From: Christine.A.Snay-at-Dartmouth.EDU (Christine A. Snay)
Date: 30 Jun 95 16:19:21 EDT
Subject:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {6304212-at-schlitzen.Dartmouth.EDU}

Subscribe
Christine A. Snay-at-dartmouth.edu




From: Michael Rock :      merock-at-u.washington.edu
Date: Fri, 30 Jun 1995 13:42:36 -0700 (PDT)
Subject: unsubscribe vacation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe merock-at-u.washington.edu