From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Thu, 29 Feb 1996 12:19:36 -0600
Subject: Re: SEM Polaroid films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

In message {Megw.230326-at-hermes} "Mount, Richard" writes:
}
} {Text_1}
} Dennis Shubitowski asked "Has anyone successfully "unstuck"
} Polaroid Type 55 negatives that have stuck together while
} drying?"

{text 2}

} Films stuck emulsion to film back seem to separate easier than those
} stuck emulsion to emulsion.
}
} ------------------------------------------------------------
} Richard J. Mount Phone: 416-813-6551
} Auditory Science Laboratory
} Department of Otolaryngology FAX: 416-813-5036
} The Hospital for Sick Children
} 555 University Ave.
} Toronto, ON, Canada M5G 1X8
}
} e-mail:richard.mount-at-mailhub.sickkids.on.ca


I agree with Richard, above, that film back stuck to emulsion is easier to
seperate, and this is the case I usually run into as I dry my Polaroid negs in
racks with the emulsion of one facing the back of the next neg.

A point that I left out in my previous note describing successful use of
overnight water soak to seperate stuck negs is how I set this up. As with many
techniques, success lies in the details of the method. I load a pair of stuck
negs into my usual neg drying rack with edges of one neg in one set of slots,
edges of other neg in adjacent set of slots. The effect is to separate the negs
where they are not stuck to enable water to get directly to the stuck area.
Success depends on getting good water soak and softening of the film emulsion.
If you were to simple lay the two stuck negs flat in a pan of water, might not
get as good a soak as when the films are seperated a bit to allow water access.

I hope this helps.




--

Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu

"MICROSCOPY & MICROANALYSIS 96" in Minneapolis, Minnesota, Aug.11-15, 1996





From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 29 Feb 1996 14:59:06 -0500 (EST)
Subject: Re: TEM;phosphate ppt in seawater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} a user brought me a sample of gelatinous unicellular algae she
} preserved in 2% phosphate buffered glutaraldehyde. Unfortunately, it was a
} marine sample, so when she fixed it, the salts in the seawater
} precipitated. She can't repeat the experiment. Any way to salvage her
} sample--i.e., separate the cells from the ppt, or get the salts back into
} soln without detroying the cells? We appreciate that the precipitation may
} have ruined her cell preservation as well, but we won't know for sure till
} we can embed and section.
}
Dear Steve,
I am not an expert in this, but FWIW, I would try rinsing in buf-
ferred saline using a lowish concentration of a buffer such as tris or
acetate. This might wash off or redissolve the external ppts. I wouldn't
think anything would get ppt out of the interiors of the cells but still
preserve the morphology. I'd pay attention to the osmolarity of the rinse,
although the 2% PO4+glut probably swelled the cells if they came directly
from seawater. Good luck.
Yours,
Bill Tivol




From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Thu, 29 Feb 1996 21:01:47 EST
Subject: Mo aperture grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

On Feb. 29, the following question was asked:

} Subject: Mo aperture grids
}
} I am looking for TEM grids with single hole of 0.8, 1.0 and 1.5 mm in
} diameter and slot of 1x2 and 0.5x2 mm. I tried several vender and did
not
} succeed. Anybody know a source of these grids?
}
We have had in production Mo support "rings" for use during the ion
milling of samples for quite some time. They are 3mm diameter OD with
an ID of 1.5 mm. Thickness is 0.5 mm. These should work out just fine
for you.

You can find them either on page 74 of the current SPI Supplies
SourceBook or else in our electronic on-line catalog at the ULR given
below. They will be found on the "page" with the ultrasonic disc
cutter.

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: South Bay Technology :      73531.1344-at-CompuServe.COM
Date: 29 Feb 96 18:47:20 EST
Subject: NEW Tripod Polisher Workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



REDUCED FEE REGISTRATION DEADLINE APRIL 30, 1996

Workshop on Tripod Polishing

Workshop Objective
This course will cover all aspects of pre-thinning and focus on final thinning
via Tripod Polishing. Due to the limited class size and the extensive hands-on
opportuinities, this course is well suited to novices as well as advanced
Tripodders. The course will include sections on:

How to do it and why should I?
What's really going on and what am I really seeing?
How to prepare small, specific area cross-sections.
The problem of wildly differing materials (eg tungsten).
Rapid preparation of TEM cross-sections.
Preparation of a wide range of materials: semiconductors, ceramics, metals,...

Hands-on Opportunity
This course will be unique in that it will provide a hands-on opportunity for
every class participant. Tripod Polishers, Polishing Wheels, and pre-thinning
equipment will be made available to all participants and actual samples will be
prepared - by the students - as part of the course. This is a great opportunity
to get your hands dirty and actually learn by doing. The instructors will walk
you through each step of the process and then let you loose on the equipment.
This course is designed to teach the Tripod Polishing technique. Silicon
samples will be provided to the students and used as the basis for the course
teaching.

Workshop Location and Dates
South Bay Technology - San Clemente, CA
Dates: Friday & Saturday - June 14-15, 1996
Friday & Saturday - October 18-19, 1996

Previous Participants (partial list)
INTEL, AMD, Motorola, LSI Logic, Conner Peripherals, Univ of Maryland, Univ of
New Mexico, UNAM (Mexico), LG Electronics (Korea), Battelle, MEMC, MVA Inc.,
Univ of Michigan, U.S. Bureau of Mines, IBM, Naval Research Lab, Purdue Univ,
Univ of Alabama, Univ of Arizona, Univ of Colorado, Univ of Wisconsin.

Class Size
Due to the intensive hands-on aspects of this course, class size will be
strictly limited to 10 participants.

Registration Fee: $795 (includes lunches and Friday night Dinner)
$695 if registration fee paid by April 30, 1996 for June
Workshop and by August 31, 1996 for October Workshop

Registration Deadline: 30 days prior to workshop

For additional Information: Monica Pflaster
South Bay Technology, Inc.
1120 Via Callejon
San Clemente, CA 92673
TEL: 800-728-2233
FAX: 714-492-1499
e-mail: sbt-at-msa.microscopy.com




Registration Form

To register for the workshop, please fill out this form and send it, with
registration fee to:

South Bay Technology, Inc.
Workshop on Tripod Polishing
1120 Via Callejon
San Clemente, CA 92673

Payment must be made in the form of a check, money order, Visa or MasterCard.
Checks must be drawn on a U.S. Bank and made payable to South Bay Technology,
Inc. Credit card orders by FAX may be sent to South Bay Technology at
714-492-1499.

Name:


Affiliation:


Address:




City: State:
Zip: Country:_________
Telephone: FAX:

e-mail:________________________

Primary sample type:




VISA MasterCard Card #_________________________________

Expiration Date________ Signature of Cardholder_________________________

Cardholder name (Please print):________________________________________

Please circle workshop you are registering for: June 14-15 October
18-19





From: Wei.Liu-Ying-at-mb.luth.se
Date: Fri, 1 Mar 1996 14:31:14 +0800
Subject: Software for SAED indexing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear microscopists,

I am looking for software (Macintosh) for SAED indexing. Does anyone know a
source of such software? Thanks for your help.
Yours sincerely,
Liuying Wei






From: peling-at-amnh.org (Peling Melville - Interdepartmental Facilities)
Date: Fri, 1 Mar 1996 11:04:26 -0500
Subject: Etching Iron Oxide?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603011445.IAA02762-at-Sparc5.Microscopy.Com}

A coworker of mine wanted to know if there was anything to etch fossils
that have been altered to iron oxide.

Thanks in advance,
Peling Melville

--------------------------------------------------------------
Peling Fong Melville
Senior Scientific Assistant
Interdepartmental Facilities
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 U.S.A.
******************************
E-mail: peling-at-amnh.org
Work #: (212) 769-5469
FAX #: (212) 769-5495






From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Fri, 1 Mar 1996 11:24:26 -0700
Subject: Ultramicrotome needed by non-profit foundation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A private non-profit foundation [401(3)(C) status] is looking for a used
ultramicrotome and glass knifemaker through a donation or inexpensive
purchase.

If you can help, please contact Dr. Gary Sibert at (505) 434-1725.

Please don't reply to this list. Thanks,

John
chandler-at-lamar.ColoState.EDU






From: lifan chen :      lfchen-at-unm.edu
Date: Fri, 1 Mar 1996 11:19:16 -0700 (MST)
Subject: Hello!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sir/Madam:

I am a microscopist in TEM Laboratory of the University of New Mexico.
I hope to get some new information from you every day. Could you send me
some infromation every day if possible?

Thank you very much!!!





From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Fri, 1 Mar 1996 12:52:10 -0700
Subject: Philips TEM available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


TEM available:

Philips CM10
Purchased 1987
With EDAX EDS system and ancillary equipment for materials lab
Currently under service contract with Philips

Contact: Lori or Frank Visconti
(303) 238-5574 (until March 21)
(303) 987-9178

Do not reply to this list. Thanks,

John
chandler-at-lamar.ColoState.EDU






From: T. Page Owen Jr :      tpowe-at-conncoll.edu
Date: Fri, 1 Mar 1996 15:24:10 -0500 (EST)
Subject: Azide and plant cell ultrastructure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603012106.PAA03222-at-Sparc5.Microscopy.Com}


Does anyone know if adding sodium azide to a plant cell suspension
culture (final concentration 0.02%) affect its ultrastructure?

It was added 30 minutes before fixing the cells in 2.5% glutaraldehyde.
Azide was used in order to stop cell activity at a specific time point.
The main interest in the ultrastructure analysis is to look at the Golgi
apparatus, cell wall, plasma membrane, and endoplasmic reticulum.

If it does (or doesn't) affect the ultrastructure; How? Which
organelles are affected the most? Are there references in the literature
about it?

Thanks in advance for your input.

Manrique Rojas
Page Owen
Department of Botany
Connecticut College





From: TKWZ51A-at-PRODIGY.COM (MR ALEXANDER W GINGELL)
Date: Fri, 01 Mar 1996 18:15:35 EST
Subject: Azide and plant cell ultrastructure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



-- [ From: Alex W. Gingell * EMC.Ver #2.10P ] --

Dear Microscopists,

I am sending this on behalf of a customer who does not have e-mail
access.

He would like to communicate with anyone who has a Tracor TN5500
analyser for sale. The customer is in Southern Ontario (Canada), so
would only consider units in Canadian or USA due to shipping costs.

Please e-mail to me directly, and I will pass the information on to my
customer.

Thanks in advance for your assistance.

Alex





From: Alan S. Pooley :      pooley-at-ahab.rutgers.edu
Date: Fri, 1 Mar 1996 10:46:11 -0500 (EST)
Subject: unsticking polaroid negatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


(snip below)

YES, virginia, polaroid negatives can usually be unstuck, just soak in
very hot water about 10 -20 minutes, then GENTLY pull them apart, GRADUALLY
turning and pulling a little untilthe whole gelatine that is stuck gets
soaked through and losened. Some times if more completely dried out, the
emulsion of one will be stuck to the second, usually the second can be saved
completely, but some of the first will be lost (natually the center, best
part of the image)
good luck

} Unsticking polaroid negatives
} that have stuck together while drying? A water soak did not
} seem to help nor did Photoflo.

} Thanks,

} Dennis Shubitowski
} University of Michigan
} School of Dentistry


Alan S. Pooley ,PhD Bivalve shell SEM & shape analysis
SEM/Morphometrics lab
Marine & Coastal Sciences
Rutgers University
908 932 8959 ext 225
Pooley-at-ahab.rutgers.edu





From: Elinor Solit :      cambrex-at-world.std.com
Date: Fri, 1 Mar 1996 11:56:32 -0500 (EST)
Subject: Re: How to prepare beryllium metal?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Yolande,

Suggest you try Paul Fischione. He is at 412-325-5444.

Best regards,

Ellie Solit

On Thu, 29 Feb 1996, Berta, Yolande wrote:

}
} Dear Microscopists,
}
} We have a chunk of beryllium, and we would like to know the thickness of the
} oxide layer that forms on the metal within a given time interval. SEM hasn't
} shown the oxide layer, so now we are thinking that maybe a TEM cross-section
} specimen would be the thing to try.
}
} Has anyone prepared beryllium for TEM before? How? Ion milling?
} Jet-polishing? How was it cut to keep the dust at a minimum?
}
} Yolande Berta
} School of Materials Science and Engineering
} 778 Atlantic Dr.
} Atlanta, GA 30332-0245
} (404)894-2545
} yolande.berta-at-mse.gatech.edu
}




From: Joe D Geller :      geller-at-world.std.com
Date: Fri, 1 Mar 1996 09:56:12 -0500 (EST)
Subject: Re: How to prepare beryllium metal?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This seems to be a perfect application for Auger electron spectroscopy
which is sensitive to the first few monolayers on the surface. The
surface is ion beam sputtered to determine the oxide depth. Auger
spectroscopy is very sensitive to Be.

If you don't have your own instrument, or one on campu, there are several
commercial laboratories providing this service, including ours.

Joe Geller
Geller MicroAnalytical Lab
426e Boston St.
Topsfield, MA 01983-1216
508 887-7000

On Thu, 29 Feb 1996, Berta, Yolande wrote:

}
} Dear Microscopists,
}
} We have a chunk of beryllium, and we would like to know the thickness of the
} oxide layer that forms on the metal within a given time interval. SEM hasn't
} shown the oxide layer, so now we are thinking that maybe a TEM cross-section
} specimen would be the thing to try.
}
} Has anyone prepared beryllium for TEM before? How? Ion milling?
} Jet-polishing? How was it cut to keep the dust at a minimum?
}
} Yolande Berta
} School of Materials Science and Engineering
} 778 Atlantic Dr.
} Atlanta, GA 30332-0245
} (404)894-2545
} yolande.berta-at-mse.gatech.edu
}




From: Nina S Allen :      nallen-at-unity.ncsu.edu
Date: Sat, 2 Mar 1996 11:31:16 -0500 (EST)
Subject: Re: Azide and plant cell ultrastructure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential

Azide inhibits the mitochondria and yes it has a profund effect on the
cytoskeleton and cytoplasmic streaming in stamen hair cells of Setcresea.
Read E. Tucker and N.S. Allen, 1986. Cell Motil and the cytoskel.
6:305-313.

On Fri, 1 Mar 1996, T. Page Owen Jr wrote:

}
} Does anyone know if adding sodium azide to a plant cell suspension
} culture (final concentration 0.02%) affect its ultrastructure?
}
} It was added 30 minutes before fixing the cells in 2.5% glutaraldehyde.
} Azide was used in order to stop cell activity at a specific time point.
} The main interest in the ultrastructure analysis is to look at the Golgi
} apparatus, cell wall, plasma membrane, and endoplasmic reticulum.
}
} If it does (or doesn't) affect the ultrastructure; How? Which
} organelles are affected the most? Are there references in the literature
} about it?
}
} Thanks in advance for your input.
}
} Manrique Rojas
} Page Owen
} Department of Botany
} Connecticut College
}
}




From: levin-at-ecsuc.ctstateu.edu (Martin Levin)
Date: Sun, 3 Mar 1996 14:28:18 -0600
Subject: Replacement Bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I received a posting asking about replacement bulbs for some piece of
equipment. Unfortunately, before I completely digested the message, I
deleted it (I guess the hand is really quicker than the eye). I have an
excellent source for bulbs of all kinds, even those that are no longer
manufactured. You might want to give them a call to see if they have what
you are looking for. They are:

Bulbs Only
954 Queen Street
Southington, CT 06489
(203) 621-0213

I might add that they are extremely helpful and ship the next day.

Martin A. Levin
Department of Biology
Eastern Connecticut State University
Willimantic, CT 06226
Phone: (860)465-4324 FAX: (860)465-5213






From: Peter Baron :      baron-at-globalnet.co.uk
Date: Sun, 3 Mar 1996 17:07:37 GMT
Subject: List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe Peter Baron






From: Sarka Lhotak :      lhotaks-at-fhs.csu.McMaster.CA
Date: Sun, 3 Mar 1996 11:48:04 -0500 (EST)
Subject: Re: Immunohistochemistry in plastic

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On 25 Feb 1996 VFN6T-at-DMT03.mcc.Virginia.EDU wrote:

} Dear Colleagues,
} I am using plastic-embedded (Lowicryl or JB-4) sections of fetal kidneys
} and would like to use standard avidin-biotin-peroxidase techniques for
} immunohistochemistry. To date, we are unable to achieve specific staining.
} We have tried two different etching protocols and a few of the JB-4 (but none
} of the Lowicryl) sections survive, but still no staining. Most references
} give little information regarding incubation times, etc. Can anyone help me} trouble-shoot or point me towards a good reference? Thanks in advance.
} Victoria F. Norwood, M.D.
} Department of Pediatrics
} University of Virginia
} vfn6t-at-dmt03.mcc.virginia.edu
}
}
Dear Victoria,
A failure to immunolabel is usually due to the alteration of the
antigenic conformation during tissue processing. Each antigen/antibody
combination requires different conditions (fixation, incubation times,
choice of resin etc.) which have to be determined by trial and error
approach. Some antibodies will never label. Though some monoclonal
antobodies label very nicely, chances are better with polyclonal antisera
(affinity purified) as they bind to several antigenic determinants of the
antigen and some might be preserved better than others.
The best reference would be the one using your particular
antibody. Others can give you only general directions for finding the
optimal protocol (e.g. Hyatt's three volumes of Immunogold labelling -I
don't have the exact reference here, sorry).
First, I would try to stain paraffin sections, or even better,
cryostat sections. If there is no staining there, chances are that it will
not work in resin. However, if this is successful, you can use it to
determine the optimal concentration and incubation time and then proceed
to Lowicryl or LR White sections (I have no experience with JB4). You
should not need to etch them. Some antigens, however, require enzyme
digestion (trypsin or pepsin). Again this you would try on paraffin or
cryosections first. You might need to experiment with fixation. For TEM, I
usually start with three fixation protocols: 4% paraformaldehyde (PF), 2%
PF+ 0.1 % glutarladehyde (GA) or 2% GA to find the optimal tissue
preservation vs. labelling conditions.
I hope this will help a bit. These methods can be very frustrating
at times. You may contact me directly at my E-mail address if you need
more details.

Sarka Lhotak
lhotaks-at-fhs.mcmaster.ca
EM Facility, McMaster University
Hamilton, Ontario





From: Ross Davidson :      davidson-at-surf.ssw.uwo.ca
Date: Sun, 3 Mar 1996 19:15:05 -0500
Subject: Heavy Metals in Animal Tissues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603040015.TAA22012-at-surf.ssw.uwo.ca}

Dear Microscopists:

We have recently been approached with an interesting analytical
problem concerning the detection of trace ( {200 ppm) levels of heavy
elements in animal
tissue. Elements of interest include selenium,nickel,arsenic,copper and many
others. The samples will be presented as thin (5 um) slices on a glass slide
substrate. Most of our work involves non-biological samples and this job
represents an interesting challenge.
We have available a Cameca 3f SIMS to carry out this work. We plan
to carry out depth profiles and also SIMS imaging.
Can anyone provide some useful tips in tackling this problem?
Also,is there any other techniques available that could also address this
question?
Any suggestions would be greatly appreciated.

Sincerely,
Ross Davidson Phone: 519-661-2173
Surface Science Western FAX: 519-661-3709
Western Science Centre, Rm G1, E-mail: davidson-at-surf.ssw.uwo.ca
University of Western Ontario Home Page: http://www.uwo.ca/ssw/
London, Ontario,
Canada N6A 5B7





From: HANS SLUIMAN :      RBG_3/HANS
Date: Sat, 2 Mar 1996 12:57:16 BST
Subject: Re: (Fwd) TEM;phosphate ppt in seawater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Forwarded message:

} a user brought me a sample of gelatinous unicellular algae she
} preserved in 2% phosphate buffered glutaraldehyde. Unfortunately, it was a
} marine sample, so when she fixed it, the salts in the seawater
} precipitated.


I would suggest trying to dissolve the precipitate by diluting the
buffer. The combination of phosphate buffer and seawater is clearly
not a good one. In the past I always used glutaraldehyde in seawater
as a primary fixative for marine algae. I suspect it's the calcium in
the seawater who is to blame.


She can't repeat the experiment. Any way to salvage her
} sample--i.e., separate the cells from the ppt, or get the salts back into
} soln without detroying the cells? We appreciate that the precipitation may
} have ruined her cell preservation as well, but we won't know for sure till
} we can embed and section.

Alternatively one may keep part of the samples and proceed with
embedding and sectioning as usual, without trying to remove the
precipitate. Sectioning may be a bit difficult but some useful
results may be possible to obtain. I wouldn't use a diamond knife in
this case, though.

Yours sincerely

Dr Stephan Helfer, SSO
Mycologist / Plant Pathologist

Royal Botanic Garden Edinburgh, Inverleith Row, EDINBURGH EH3 5LR,
Scotland UK

http://www.rbge.org.uk

phone: +44 (0)131 552 7171 ext 280
or +44 (0)131 459 0446-280 (direct digital VoiceMail line)
fax: +44 (0)131 552 0382
============================
1896
* BRITISH
* MYCOLOGICAL
* SOCIETY
1996
A century of fungal science
============================




From: knecht-at-uconnvm.uconn.edu (David Knecht)
Date: Mon, 4 Mar 1996 08:20:26 -0600
Subject: Hanging Drop slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can someone point me to some information about how to use a hanging drop
slide. This should be obvious I guess, but I am sure there are some little
tricks that we would rather not have to figure out the hard way. Thanks-
Dave

Dr. David Knecht
Department of Molecular and Cell Biology
University of Connecticut
U-125
Storrs, CT 06269
Knecht-at-uconnvm.uconn.edu






From: Dwight Beebe :      beebed-at-ere.umontreal.ca
Date: Mon, 4 Mar 1996 10:11:04 -0500 (EST)
Subject: Affinity-purified Ab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good morning,
I've recently been trying to do immunolocalization at the EM
level with affinity-purified Ab. My problem is that the purified Ab's
are in an ammonium thiocyanate solution that chews up both the Ni grids
and the sections (LR White). I know it's the solution because the
polyclonal serum immunos are fine. Does anyone with experience purifying
Ab's and using them for TEM immuno care to chime in with some pertinent
advice? I'd appreciate hearing any and all suggestions. Many thanks!

Dwight Beebe
Institut de recherche en biologie vegetale beebed-at-ere.umontreal.ca
Dept. de sciences biologiques Voice:514-872-4563
Universite de Montreal FAX:514-872-8496
4101, rue Sherbrooke est
Montreal, PQ H1X 2B2 Canada





From: kennedy-at-nsi.edu (grace kennedy)
Date: Mon, 4 Mar 1996 08:28:45 -0800
Subject: plastic immuno

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: grace-at-popmail
Message-Id: {ad60c84b00021004ef08-at-[198.147.151.19]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Re: Immunostaining of plastic (JB-4, LRWHite) sections. According to the
circular accompanying the LR White resin, avidin-biotin binding is masked
by the plastic--they recommend using a PAP technique instead. The same may
be true with the methacrylate resins too, so your problem could conceivably
be caused by your choice of embedding medium. Needless to say, the advice
to examine other aspects of the system is of course wise especially if
you've never tried this specific detection using another preparation
technique. I found it to be amazing what antigens can really withstand
once the plastic masking problem has been solved.... Good luck. Grace
Kennedy






From: Liang, Long :      LLIANG-at-is.arco.com
Date: 04 Mar 1996 09:56:09 CST
Subject: EPMA -- N in graphite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MACMS.LLIANG.584314100096064FMACMS-at-IS.ARCO.COM}

Dear Microscopists,

I am trying to analyze graphite grains that are scattered in a sample
for their nitrogen contents. The mass absorption coefficient for
nitrogen Ka line in the carbon matrix is very very high (23,586).
Moreover, the nitrogen is a minor or trace element in the graphite.

Is this an impossible task for the probe analysis ? Are there any other
techniques can to used to complete this task ? Thanks.

Long Liang
ARCO EPMA/SEM Lab







From: CASORIA-at-FOZZIE.SSEC.HONEYWELL.COM
Date: Mon, 4 Mar 1996 12:27:59 -0600 (CST)
Subject: Benchmarking other semiconductor labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fellow Microscopists,
We are trying to locate other analytical laboratories that might be interested
in benchmarking. Specifically, we would like to benchmark types of services
offered, size, education and expertise of staff, cycle-time, and SPC.
We are an semiconductor facility but may find acceptable benchmarks with labs
that provide .services to other than semiconductor houses.
Thanks!
Please reply to:
Carolyn Casoria
Honeywell/SSEC
-at-
casoria-at-ssec.honeywell.com




From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Mon, 4 Mar 1996 11:19:08 -0400 (EDT)
Subject: Hitachi 200kV TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English


Due to a consolidation of two microscopy facilities, we would like
to reluctantly part with the following excess instrument which is in
excellent condition and under service contract:


HITACHI H800/810 200 kV TEM
LaB6 electron gun
Motorized goniometer stage
SEM/STEM capabilities
Tracor/Northern (Noran) 5402 energy dispersive X-ray microanalysis
system with a 8502/S Advanced Image Analysis system
Special specimen holder for X-ray analysis of sectioned specimens
Special specimen holder for X-ray analysis of bulk specimens
Rotation/tilt specimen holder for grids.

Shipping the instrument is buyer's responsibility.

BONUS: If the purchase is completed before May 15th, 1996, a
HITACHI HHS-2R SEM (in storage) will be offered free of
charge to the H800 buyer.

Please contact me directly if interested.

*************************************************************************
M.V. Parthasarathy
Professor of Plant Biology &
Director, Cornell Integrated Microscopy Center
Section of Plant Biology
228 Plant Science Building
Cornell University, Ithaca, NY 14850
Telephone: (607) 255-1734
Fax: (607) 255-5407
E-mail: mvp2-at-cornell.edu
***********************************************************************




From: SNYDER, JOSEPH :      JTS-at-rrccorp.mhs.compuserve.com
Date: 04 Mar 96 14:27:24 EST
Subject: Wrong Subscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have inadvertently given the wrong email address during my subscription
attempt.

Please change: JTS-at-rrc.mhs.compuserve.com

to: JTS-at-rrccorp.mhs.compuserve.com

You have my sincere apologies if this has caused you any problems.

Joe Snyder






From: SNYDER, JOSEPH :      JTS-at-rrccorp.mhs.compuserve.com
Date: 04 Mar 96 15:12:32 EST
Subject: Desubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please desubscribe: JTS-at-rrc.mhs.compuserve.com






From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Mon, 4 Mar 1996 09:09:09 GMT
Subject: Re: Digest

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Is a digest now available for the Microscopy list? Thanks for any information.
}
}
} Blystone in Texas
}
} **********************
} ROBERT V. BLYSTONE PHONE:(210)736-7243
} DEPARTMENT OF BIOLOGY FAX:(210)736-7229
} Trinity University E-Mail:RBlyston-at-Trinity.edu
} 715 Stadium Drive
} San Antonio, TX, 78212
}
} } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } }

We currently archive items of interest to biologists. It is not yet
"searchable" but will be soon. It can be found at :

http://www.biotech.ufl.edu/~emcl/tips.html
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-





From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 4 Mar 1996 14:02:45 EST
Subject: Re: Hanging Drop slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: Dr. David Knecht

To make a hanging drop slide: You will need a depression slide, a
square coverslip, some petrolatum, and the liquid suspension of what
you wish to view.

Place a small spot of petrolatum on each of the 4 corners of the
coverslip. Place a drop of your suspension in the center of the
coverslip. Invert a depression slide over the drop, allowing the
petrolatum to attach the coverslip to the depression slide. Quickly
(but carefully) invert the slide so that the coverslip is oriented
"up", and the drop is hanging into the slide depression.

That's all there is to it.

W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 4 Mar 1996 14:02:29 EST
Subject: Re: Hanging Drop slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: Dr. David Knecht

To make a hanging drop slide: You will need a depression slide, a
square coverslip, some petrolatum, and the liquid suspension of what
you wish to view.

Place a small spot of petrolatum on each of the 4 corners of the
coverslip. Place a drop of your suspension in the center of the
coverslip. Invert a depression slide over the drop, allowing the
petrolatum to attach the coverslip to the depression slide. Quickly
(but carefully) invert the slide so that the coverslip is oriented
"up", and the drop is hanging into the slide depression.

That's all there is to it.

W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: Beverly E Maleeff
Date: 4 Mar 96 11:53:08 EDT
Subject: March PSM Meeting Notice

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9603042108.AA1032-at-pho903.sbphrd.com}
To: microscopy {microscopy-at-Sparc5.Microscopy.Com}
{Beverly_E_Maleeff-at-sbphrd.com}

Philadelphia Society for Microscopy (PSM) March Meeting

When: Tuesday, March 12, 1996

Where: Laboratory for the Research of Science and Materials (LRSM),
on the campus of the University of Pennsylvania, 33rd and Walnut Sts.,
Philadelphia, PA

Program: 5:30 PM Social hour
6:30 PM Dinner Members $12.00, Non-members $15.00
7:30 PM Speaker

Speaker: Dr. Roberto F. Nicosia, Dept. of Pathology,
Medical College of Pennsylvania and Hahnemann University
Philadelphia, PA

"The Cell Biology of Blood Vessel Growth: Light microscopic,
immunohistochemical and ultrastructural studies"

Reservations will be taken by Ms. Pat Overend at the University of Pennsylvania,
215/898-8337. Deadline for dinner reservations is Friday, March 8.
Cancellations
must be received no later than 5:00 PM, March 11, 1996.

If you have any questions regarding the meeting please contact Rollin Lakis at
the
University of Pennsylvania, 215/898-8718.






From: VOS-at-CSSS.la.asu.edu
Date: Mon, 4 Mar 1996 17:42:53 -0700 (MST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please unsubscribe. Thank you




From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 4 Mar 1996 14:02:36 EST
Subject: Re: Hanging Drop slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: Dr. David Knecht

To make a hanging drop slide: You will need a depression slide, a
square coverslip, some petrolatum, and the liquid suspension of what
you wish to view.

Place a small spot of petrolatum on each of the 4 corners of the
coverslip. Place a drop of your suspension in the center of the
coverslip. Invert a depression slide over the drop, allowing the
petrolatum to attach the coverslip to the depression slide. Quickly
(but carefully) invert the slide so that the coverslip is oriented
"up", and the drop is hanging into the slide depression.

That's all there is to it.

W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Mon, 4 Mar 1996 19:31:07 -0600
Subject: Re: Air Dried Bacteria?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02120d00ad6148d6ed4b-at-[128.174.107.186]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I've had excellent results air drying from hexamethylsilizane at
room temperature. Peldri will also work if you've got any left. Filtering
through a filter works fine.
But what % size change are you looking for? Recall the measuring
things, especially biological specimens in the SEM is dicey at best--and
can be grossly misleading.
Phil Oshel

} Hi All:
}
} I have been asked to examine, using the SEM, some bacteria that have been
} exposed to a drug solution. The interest is in whether the exposure to the
} drug solution has altered the bacterial shape and size (mostly size) and
} thus effect its retention on a filter. I propose to simply filter the
} bacterial suspension onto a polycarbonate membrane, air dry, mount, sputter
} coat, and examine.
} Damian Neuberger
} Email: neuberd-at-baxter.com

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 4 Mar 1996 14:03:03 EST
Subject: Re: Hanging Drop slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: Dr. David Knecht

To make a hanging drop slide: You will need a depression slide, a
square coverslip, some petrolatum, and the liquid suspension of what
you wish to view.

Place a small spot of petrolatum on each of the 4 corners of the
coverslip. Place a drop of your suspension in the center of the
coverslip. Invert a depression slide over the drop, allowing the
petrolatum to attach the coverslip to the depression slide. Quickly
(but carefully) invert the slide so that the coverslip is oriented
"up", and the drop is hanging into the slide depression.

That's all there is to it.

W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 4 Mar 1996 14:03:07 EST
Subject: Re: Hanging Drop slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: Dr. David Knecht

To make a hanging drop slide: You will need a depression slide, a
square coverslip, some petrolatum, and the liquid suspension of what
you wish to view.

Place a small spot of petrolatum on each of the 4 corners of the
coverslip. Place a drop of your suspension in the center of the
coverslip. Invert a depression slide over the drop, allowing the
petrolatum to attach the coverslip to the depression slide. Quickly
(but carefully) invert the slide so that the coverslip is oriented
"up", and the drop is hanging into the slide depression.

That's all there is to it.

W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: GREG VAUGHN MARTIN :      gmartin-at-welchlink.welch.jhu.edu
Date: Mon, 4 Mar 1996 13:53:53 -0500 (EST)
Subject: Re: Affinity-purified Ab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dwight --
Try dialyzing the affi-pure Ab into Tris-Buffered Saline. I use
10mM Tris, 500mM NaCl, pH 7.2 for use on LRGold sections. Avoid
phosphate as it causes lead stain pption. Good Luck,

Greg Martin
Dept. of Cell Biology and Anatomy
Johns Hopkins School of Medicine

On Mon, 4 Mar 1996, Dwight Beebe wrote:

} Good morning,
} I've recently been trying to do immunolocalization at the EM
} level with affinity-purified Ab. My problem is that the purified Ab's
} are in an ammonium thiocyanate solution that chews up both the Ni grids
} and the sections (LR White). I know it's the solution because the
} polyclonal serum immunos are fine. Does anyone with experience purifying
} Ab's and using them for TEM immuno care to chime in with some pertinent
} advice? I'd appreciate hearing any and all suggestions. Many thanks!
}
} Dwight Beebe
} Institut de recherche en biologie vegetale beebed-at-ere.umontreal.ca
} Dept. de sciences biologiques Voice:514-872-4563
} Universite de Montreal FAX:514-872-8496
} 4101, rue Sherbrooke est
} Montreal, PQ H1X 2B2 Canada
}
}




From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 4 Mar 1996 13:45:26 EST
Subject: Re: Affinity-purified Ab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To Dwight Beebe:

Antisera can be easily separated from undesirable salts by
dialysis...this is by far the simplest method. Obtain a short length
of dialysis tubing, soak it in water or buffer which will allow it to
be opened up into a tube. Tie off one end, fill with your antibody
solution, then tie off the other end. Immerse it into whatever
buffer you want to antibody to end up in (I suggest either TRIS
buffered or phosphate buffered saline. Use 1000X as much buffer as
antiserum, ie dialyse 1 cc of antibody against 1 liter of water for
24 hours, then change the buffer and continue for another 24 hours.
I recommend the last buffer change to contain 0.001 mM sodium azide
to retard microbial growth.

This will hopefully solve your problem.
W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: rw9-at-psu.edu (Rosemary A. Walsh)
Date: Tue, 5 Mar 1996 00:18:25 -0500
Subject: Re: Air Dried Bacteria?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"'Microscopy '" {Microscopy-at-Sparc5.Microscopy.Com}

At 3:57 PM 3/4/96 -0800, Neuberger, Damian wrote:
} Hi All:
}
} I have been asked to examine, using the SEM, some bacteria that have been
} exposed to a drug solution. The interest is in whether the exposure to the
} drug solution has altered the bacterial shape and size (mostly size) and
} thus effect its retention on a filter. I propose to simply filter the
} bacterial suspension onto a polycarbonate membrane, air dry, mount, sputter
} coat, and examine. This will supposedly work for the Gram positive microbes
} but may not work for the Gram negatives (they are said to have a much
} thinner wall and could shrink). An alternate approach is to heat fix the G-
} to a glass coverslip, sputter coat, and examine.
}
} Is this approach OK or is there a better way to do this, perhaps using a
} fixative? I would rather not have to go the route of CPD. Any suggestions
} will be greatly appreciated!!!
}
} Damian Neuberger
} Email: neuberd-at-baxter.com

Dear Damian,
One alternative to CPD is to filter, fix, dehydrate and follow the
final 100% ETOH dehydration with 3 baths of 100% HMDS
(Hexamethyldisilizane). Polycarbonate membranes handle the solvent very
well. They dry in minutes and you can quickly afix them to stubs and
sputter coat.
Rosemary WAlsh, EM Facility, PSU






From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Mon, 4 Mar 1996 15:33:16 -1000 (HST)
Subject: Pt/Ir coating for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In the absence of a means to coat biological specimens with Cr for
high(er) resolution SEM, we use the electron guns in our Balzers 400
freeze-fracture device to coat with various metals and carbon. Several
years ago Hans Ris suggested Pt/Ir/C coating, and we have had very good
results with this, although we have to constantly struggle with some
of the parameters. I am interested in hearing from others who use
this technique to see if we can compare notes. For example, should the
coating be done cold or at room temperature? We have had some success
under both conditions. What kinds of distances, angles, vacuums, etc.?

On a related note, we had been purchasing 80% Pt/20% Ir wire
and balling it up and melting it into drilled out C rods (supplied by
Bal-Tec), but this was a pain to do as compared with using the Pt
inserts made for the drilled rods. Bal-Tec would make the Pt/Ir pellets
to order, but at substantial cost. Larry Stevens at Stevens
Metallurgical Corp. has agreed to make the Pt/Ir pellets and/or rods at
a very reasonable price if there appears to be enough of an interest. I
would be like to hear from others who might buy Pt/Ir pellets so that we
might encourage Mr. Stevens to do so!

Aloha from sunny, warm Hawaii (and surf's up),
Tina
*****************************************
Tina (Weatherby) Carvalho *
Biological Electron Microscope Facility *
University of Hawaii *
(808) 956-6251 *
tina-at-ahi.pbrc.hawaii.edu *
http://www.pbrc.hawaii.edu/bemf/ *
*****************************************





From: Neuberger, Damian :      neuberd-at-engrnd.roundlake.baxter.com
Date: Tue, 5 Mar 1996 07:56:02 GMT+0200
Subject: Air Dried Bacteria?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"Neuberger, Damian" {neuberd-at-engrnd.roundlake.baxter.com}



Hi All:

I have been asked to examine, using the SEM, some bacteria that have been
exposed to a drug solution. The interest is in whether the exposure to the
drug solution has altered the bacterial shape and size (mostly size) and
thus effect its retention on a filter. I propose to simply filter the
bacterial suspension onto a polycarbonate membrane, air dry, mount, sputter
coat, and examine. This will supposedly work for the Gram positive microbes
but may not work for the Gram negatives (they are said to have a much
thinner wall and could shrink). An alternate approach is to heat fix the G-
to a glass coverslip, sputter coat, and examine.

Is this approach OK or is there a better way to do this, perhaps using a
fixative? I would rather not have to go the route of CPD. Any suggestions
will be greatly appreciated!!!

Damian Neuberger
Email: neuberd-at-baxter.com

Hello Damian!

No fix and air drying will almost certainly change the shape and
size of the bacteria. If you wish to avoid critical point drying
you could try fixation followed by a CPD substitute such as
HMDS (hexamethyldisilazane), TMS (tetramethylsilane) or Peldri
(if available). Alternatively, freeze-drying, cryo-SEM or LVSEM
would be preferable to air-drying without fixation.

Regards,



Robin Cross
Director : EM Unit, Rhodes University, Grahamstown, South Africa
eurc-at-giraffe.ru.ac.za (tel: +27 461 318168 - fax: +27 461 24377)




From: CSEDAX-at-ARCRIDE.EDU.AR
Date: Tue, 5 Mar 1996 09:28 -0300
Subject: X-Ray Fluorescence: listserver and sources for buying supplies needed.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear microscopists,

besides SEM and SEM-EDS analysis we are doing XRF
services since not so long ago. We know this may not be the forum to asking
about XRF but, we have experienced the kindness of many of you out there.

Our questions are:

* do any of you know of any listserver for X-Ray Fluorescence Spectroscopists
to discuss ideas with?

* we're desperately needing mylar or polypropilene film for doing XRF
analysis, do you know a source for XRF supplies ? We'd appreciate very
much receiving postal or e-mail addresses or fax to contact them as soon
as possible.

We thank all of you for your attention,

Angel Di Giandomenico
cnono-at-arcride.edu.ar

Centre for Research and Development
Santa Fe - Argentina




From: James S MArtin :      James.S.Martin-at-williams.edu
Date: Tue, 5 Mar 1996 07:41:04 -0500 (EST)
Subject: field trips

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


As part of an introductory course on microscopy, I am trying to set up
field trips so students can observe practical applications of acoustic
and scanning tunneling microscopy.

So ... I am looking for industrial, commercial or academic labs within
driving distance of Williamstown, MA that would be willing to spend some
time with a small group of eager undergraduates. Dates to be determined.

Thanks.

James Martin
Williams College




From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Tue, 5 Mar 1996 09:03:58 GMT
Subject: Re: Air Dried Bacteria?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} To: "Neuberger, Damian" {neuberd-at-engrnd.roundlake.baxter.com}
} From: gwe-at-biotech.ufl.edu (Greg Erdos)
} Subject: Re: Air Dried Bacteria?
} Cc:
} Bcc:
} X-Attachments:
}
} }
} } Hi All:
} }
} } I have been asked to examine, using the SEM, some bacteria that have been
} } exposed to a drug solution. The interest is in whether the exposure to the
} } drug solution has altered the bacterial shape and size (mostly size) and
} } thus effect its retention on a filter. I propose to simply filter the
} } bacterial suspension onto a polycarbonate membrane, air dry, mount, sputter
} } coat, and examine. This will supposedly work for the Gram positive microbes
} } but may not work for the Gram negatives (they are said to have a much
} } thinner wall and could shrink). An alternate approach is to heat fix the G-
} } to a glass coverslip, sputter coat, and examine.
} }
} } Is this approach OK or is there a better way to do this, perhaps using a
} } fixative? I would rather not have to go the route of CPD. Any suggestions
} } will be greatly appreciated!!!
} }
} } Damian Neuberger
} } Email: neuberd-at-baxter.com
} }
} } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } }
}
} Get your bacteria to attach to a polylysine coated cover slip. Fix with
glut and osmium, dehydrate and dry via HMDS (two changes of
hexamethyldisilazane after 100% alcohol, then air dry finishing up with ten
minutes in a 60 C oven). Then sputter coat Will give you much better
results than heat fixing and air drying.
}
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-





From: thsu-at-MSC.nthu.edu.tw (Tung Hsu) (by way of chandler-at-lamar.Colostate.EDU (John
Date: Tue, 5 Mar 1996 09:04:58 -0700
Subject: Air Dried Bacteria

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Chandler))

WANT TO BUY OR ACCEPT DONATION:
electron microscope

Usable JEOL JEM-100C, 100CX, 100S , or equivalent.
Must be in OK running condition.
Accessories, high resolution, etc. unimportant.
Will pay for shipping.
This is for my colleague studying polymer materials.
You may contact me if you have one to give away or have any lead.
Thank you in advance.

Tung Hsu
Material Science Center
National Tsing-hua University
Hsin-chu 30043
TAIWAN

telephone: 886-35-724951 (with answering machine)
886-35-715131 ext 5393
fax: 886-35-713208
E-mail: thsu-at-msc.nthu.edu.tw






From: Cheri Owen :      cowen-at-cmb.biosci.wayne.edu
Date: Tue, 5 Mar 1996 12:29:18 -0500 (EST)
Subject: Used TEM for sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Recently, I sent a message to you all about a used TEM, to see if there
was any interest. I have had some, so here is the real offer.

We have a Philips 201c TEM for sale, it is approximately 20 years old.
It is functional, but lack of use has caused the vacuum to be less than
optimal. With a little TLC this scope could be an excellent workhorse.
The investigators who own the scope would be willing to sell it for
parts, but they want it removed from their facility, not just scavenged,
as they need to make room for a new scope.

Thanks in advance

Cheri Owen
Wayne State University
Detroit Neurotrauma Institute
Detroit, MI
(313)577-4648




From: Cheri Owen :      cowen-at-cmb.biosci.wayne.edu
Date: Tue, 5 Mar 1996 12:29:18 -0500 (EST)
Subject: Used TEM for sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Recently, I sent a message to you all about a used TEM, to see if there
was any interest. I have had some, so here is the real offer.

We have a Philips 201c TEM for sale, it is approximately 20 years old.
It is functional, but lack of use has caused the vacuum to be less than
optimal. With a little TLC this scope could be an excellent workhorse.
The investigators who own the scope would be willing to sell it for
parts, but they want it removed from their facility, not just scavenged,
as they need to make room for a new scope.

Thanks in advance

Cheri Owen
Wayne State University
Detroit Neurotrauma Institute
Detroit, MI
(313)577-4648




From: Ian MacLaren :      MACLARIZ-at-novell2.bham.ac.uk
Date: 5 Mar 1996 16:14:27
Subject: TEM: preparation of ceramic fibres

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: Microscopy-at-Sparc5.Microscopy.Com

Dear all,
Has anyone out there prepared non-conducting ceramic fibres for TEM, either
across or along axis? If so, how did you do it?
We have some mullite fibres we would like to look at and although we have
some ideas about how to prepare specimens we would appreciate any ideas or
feedback on this topic.

Thanks

Dr. Ian MacLaren,
IRC in Materials for High Telephone: 0121 414 3447
Performance Applications, FAX: 0121 414 3441
The University of Birmingham, email: I.MacLaren-at-bham.ac.uk
Birmingham B15 2TT,
England




From: Joe D Geller :      geller-at-world.std.com
Date: Tue, 5 Mar 1996 11:56:41 -0500 (EST)
Subject: Re: EPMA -- N in graphite

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nitrogen in graphite can be detected to a concentration level of
approximately 0.08% in a relatively short amount of time using the LDE
type of metal crystal in a wavelength dispersive x-ray spectrometer
mounted on a electron probe microanalyzer. The spatial resolution would
be on the order of 1 or 2 micrometers.

If you don't have this type of instrumentation there are commercial
laboratories offering this service, including ours.

Joe Geller
Geller MicroAnalytical Laboratory
426e Boston ST.
Topsfield, MA 01983-1216
508 887-7000 fax 887-6671
On 4 Mar 1996, Liang, Long wrote:

} Dear Microscopists,
}
} I am trying to analyze graphite grains that are scattered in a sample
} for their nitrogen contents. The mass absorption coefficient for
} nitrogen Ka line in the carbon matrix is very very high (23,586).
} Moreover, the nitrogen is a minor or trace element in the graphite.
}
} Is this an impossible task for the probe analysis ? Are there any other
} techniques can to used to complete this task ? Thanks.
}
} Long Liang
} ARCO EPMA/SEM Lab
}
}
}




From: Mail Delivery Subsystem :      MAILER-DAEMON-at-LaserMaster.Com
Date: Tue, 5 Mar 1996 10:45:31 -0600 (CST)
Subject: Returned mail: User unknown

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Normal
boundary="KAA11766.826044331/spica.LaserMaster.Com"
Auto-Submitted: auto-generated (failure)





From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Tue, 5 Mar 1996 17:30:19 -0600
Subject: LM: lab manuals, textbooks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130505ad627e292b63-at-[131.230.97.68]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I am looking for good reference books and lab manuals containing protocols
for use in a course on histotechnique\microtechnique. Anyone have some good
recent and classical books (Title, Author, Publishers)? Many thanks.

#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: LSFY69A-at-PRODIGY.COM (MR KARL H BERGER)
Date: Mon, 04 Mar 1996 20:21:53 EST
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


LSFY69A -at- prodigy.com





From: wschweitzer-at-access.ch (Wolf Schweitzer)
Date: Wed, 6 Mar 1996 02:25:07 GMT
Subject: Re: external computer drives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


concerning SyQuest Compatibility:

I am using an external SyQuest 44MB Drive:

- no problems on MacPlus, SE/30, IIcx, Quadra 700/800/900, PowerPC 8500
- STOPS MACHINE from booting on Quadra 900 with Apple PPC upgrade card !!

Greetings from Switzerland, Wolf Schweitzer.



----------------------------------------------
Wolf Schweitzer, MD - CH 8596 Scherzingen
E-Mail wschweitzer-at-access.ch
http://www.access.ch/whoiswho/wschweitzer.html






From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Tue, 5 Mar 1996 20:08:19 -0600
Subject: Re: LM: lab manuals, textbooks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02120d00ad62a295d861-at-[128.174.107.222]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

John,
Best books I know of are John Kiernan's _Histological and
Histochemical Methods_ and the Biological Stain Commission's _Staining
Procedures_ (9th or 10th [or...?] edition). I don't have Kiernan's book to
hand, but if he doesn't respond, I can send you the correct info. (Same for
_Staining Procedures_.) I heard there was a good EM book with some
procedures written by some guy in southern Illinois...
Phil Oshel

} I am looking for good reference books and lab manuals containing protocols
} for use in a course on histotechnique\microtechnique. Anyone have some good
} recent and classical books (Title, Author, Publishers)? Many thanks.
}
} #############################################################################
} John J. Bozzola, Ph.D., Director
} Center for Electron Microscopy
} Southern Illinois University
} Carbondale, IL 62901-4402
} U.S.A.
} Phone: 618-453-3730
} Fax: 618-453-2665
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/departments/shops/cem.html
} #############################################################################

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: Dirk.Voeste-at-rz.ruhr-uni-bochum.de
Date: Wed, 6 Mar 1996 17:07:41 +0000
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sirs,

please subscribe,

Dr. D. Voeste




From: Andrews Lab
Date: 3/5/96 6:17 PM
Subject: >Question about HVEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1386012229.51338-at-QuickMail.Yale.edu}



--------------------------------------
Do I have to caot the grids ?
Thanking you in advance for the suggestions, i am looking forward to try this
technique
annette Bakker







From: Daniel L. Callahan :      dlc-at-owlnet.rice.edu
Date: Wed, 6 Mar 1996 11:16:09 -0600 (CST)
Subject: silicon transmission

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603061724.MAA18470-at-thomas.ge.com}


Good Day:

Is there an available color-spectrum illustrating the color changes
undergone in silicon as it is thinned. This effect is routinely used for
final polishing but I have never actually seen a quanitative correlation
for just how thin 'yellow' transmitting silicon is, for example. I
presume it is also orientation dependent, but don't know to what degree.

A color-guide would make a very useful HTML page for new users.

Daniel L. Callahan
Department of Mechanical Engg. and Materials Science
Rice University
dlc-at-owlnet.rice.edu




From: Secretaire-at-sc.ucl.ac.be
Date: Wed, 6 Mar 1996 18:19:51 +0100
Subject: LM audio visual support

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good morning
I am looking for audio visual aids (video, CDŠ) for illustrating
theory and/or possibilities of all kinds of LM. This is for use in a
course on investigation methods. Recent reference books or booklets edited
by manufacturers are also welcome.
I'd appreciate your suggestions. Many thanks!

Prof. Jean-Pierre Auquiere
Universite catholique de Louvain (UCL)
Secr. Acad. Faculte des Sciences Unité de botanique generale (BOTA)
Place des Sciences, 2 Place Croix du Sud, 4
B-1348 Louvain la Neuve, Belgium B-1348 Louvain la Neuve, Belgium
tel + 32 10 47 86 78 tel + 32 2 764 51 22
fax + 32 10 47 28 37 fax + 32 2 764 72 55
E mail secretaire-at-sc.ucl.ac.be E mail auquiere-at-bota.ucl.ac.be






From: Doug Keene :      DRK-at-SHCC.ORG
Date: Wed, 06 Mar 1996 14:22:10 -0800 (PST)
Subject: photographic processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Considering that photography is still so important to our profession, it is
likely that in this group are a few critical individuals with personal
experiencee using dry-to-dry black and white photographic processors. Would
anyone care to recommend a counter-top processor that they are happy with? I
wouldn't object to knowing which ones to stay away from either.

Many thanks,

Doug Keene
Shriners Hospital for Crippled Children
Portland, Oregon




From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Wed, 6 Mar 1996 08:53:30 -0800 (PST)
Subject: Re: LM: lab manuals, textbooks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: glenmac-at-homer09.u.washington.edu


A useful introductory text on LM sample processing and staining methods is
Animal Tissue Techniques,
Gretchen L. Humason, W.H. Freeman, I believe the 3 rd edition is the
most recent.

Regards,

Glen MacDonald
Research Scientist
Hearing Research Laboratories of the
Virginia Merrill Bloedel Hearing Research Center
Box 35-7923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156
glenmac-at-u.washington.edu


On Tue, 5 Mar 1996, John. J. Bozzola wrote:

} I am looking for good reference books and lab manuals containing protocols
} for use in a course on histotechnique\microtechnique. Anyone have some good
} recent and classical books (Title, Author, Publishers)? Many thanks.
}
} #############################################################################
} John J. Bozzola, Ph.D., Director
} Center for Electron Microscopy
} Southern Illinois University
} Carbondale, IL 62901-4402
} U.S.A.
} Phone: 618-453-3730
} Fax: 618-453-2665
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/departments/shops/cem.html
} #############################################################################
}
}
}




From: zaluzec-at-Microscopy.Com (Nestor J. Zaluzec)
Date: Wed, 6 Mar 1996 17:53:32 -0600
Subject: MSA/MAS/MSC-SMC Comprehensive Info Location

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



If you would like to access the most up to date information
about the joint meeting between MSA, MAS & MSC-SMC
you can get everything on-line on the WWW.

This includes:
Scientific Program
Short Courses
Posters
Special Awards
Financial Support
Social Events
How to submit manuscripts
Registration Forms
Dates, Times, Deadlines, etc...

Just access the MSA WWW site at

http://www.msa.microscopy.com

Look for the hot links to "Registration & Meeting Info - 1996s"



Cheers... Nestor

Your Friendly Neighborhood SysOp
&
the Microscopy & Microanalysis Program Chair for 1996






From: William Tivol :      tivol-at-wadsworth.org
Date: Wed, 6 Mar 1996 12:09:03 -0500 (EST)
Subject: Re: TEM: preparation of ceramic fibres

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Has anyone out there prepared non-conducting ceramic fibres for TEM, either
} across or along axis? If so, how did you do it?

Dear Ian,
This may or may not be applicable, but with the high-voltage EM
(lower interaction cross-section) and very low beam currents, we have been
able to image non- or poorly-conducting specimens without charging or heat-
ing problems. If this holds true for your instrument (e.g. you have an
IVEM), the advantage is that there may be no preparation required; the
disadvantage is that with these low doses you will either need a very sen-
sitive recording medium--LoDose or other x-ray film, image plates or in-
tensified or slow-scan CCD--or long exposures. The large grain or speci-
men drift could make the images useless. Good luck.
Yours,
Bill Tivol




From: sbarlow-at-sunstroke.sdsu.edu (Steve Barlow)
Date: Wed, 6 Mar 1996 10:44:27 -0800
Subject: Call for papers, 4th Annual Microscopy Colloquium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02110100ad638b865bb5-at-[130.191.238.144]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

4th ANNUAL MICROSCOPY COLLOQUIUM

held jointly with:

California State University
and the
California Society For Microscopy

May 4 & 5, 1996
San Diego State University

REGISTRATION AND ABSTRACTS

The abstracts will be compiled into the "Proceedings of the Fourth Annual
Microscopy Colloquium" and will be published in Microscopy Research and
Technique. Abstract information and details on email submission of
abstracts available on request.

Abstracts must be received by March 29,1996

Registration fees (by March 29):
$35/regular, $10/student, $50/vendor

Late Registration fees (March 29):
$45/regular, $10/student, and $60/vendor.

Please send the enclosed registration form whether or not you intend to
make a presention

PLATFORM AND POSTER PRESENTATIONS

Papers discussing any aspects of microscopy in cell, developmental,
and structural biology, or material and geological sciences are welcome.
Microscopy techniques papers and student presentations are particularly
encouraged.
Platform presentations will be held on Saturday, May 4th and Sunday
morning, May 5th. Posters will be displayed Saturday, May 4, with authors
present.

PROGRAM

Friday - May 3:
4:30 p.m. Business meeting for CSU delegates at Comfort Inn, La Mesa

Saturday - May 4:
8:00 a.m. Registration/Aztec Hall
San Diego State University
9:00 a.m. Opening remarks
9:15 a.m. Platform presentations
10:30 a.m. Coffee break
10:45 a.m. Platform presentations
12:00 p.m. Buffet lunch
1:30 p.m. Platform presentations
3:00 p.m. Coffee break
3:15 p.m Poster presentations
4:00 p.m. Poster authors present
4:45 p.m. Reception/ Banana Quad
Tour of the SDSU EM Facility
6:00 p.m. Banquet
7:30 p.m. Keynote address

Sunday-May 5
9:00 a.m. Platform presentations
10:45 a.m. Coffee break
11:00 a.m. Panel discussion
"Remote Access To Instrumentation"
12:00 p.m. Closing remarks



Registration Form
CSU Microscopy Colloquium
May 4 & 5, 1996
San Diego

_____________________________________________________
Name

_____________________________________________________
Address

_____________________________________________________
City State Zip

______________________ ________________________
Phone e-mail

Presentation:
O Platform O Poster O No Presentation

Meal Preference:
O Chicken O Fish O Vegetarian

Make checks payable to "EM Facility/Colloquium"

SEND TO:

Dr. Steven Barlow
EM Facility/Biology Dept.
San Diego State University
5500 Campanile Drive
San Diego, CA. 92182-4614


(619) 594- 4523
sbarlow-at-sunstroke.sdsu.edu

HOUSING:

The Comfort Inn, La Mesa

2 miles from SDSU

Housing for the Colloquium will be available at the Comfort Inn, 8000
Parkway Drive, La Mesa. The room rate is $42.00 per night, plus tax. You
must make your own reservations. Please call (619) 698-7747 and specify
that you will be attending the Microscopy Colloquium.

Shuttle Service:

Airport shuttle service is available from several companies. We
recommend the California Sunshine Shuttle Service at $12 per person to SDSU
and the Comfort Inn. Please make advanced reservations by calling collect
(619) 443-7900, 6 a.m. to 9 p.m. PST and specify that you will be attending
the Microscopy Colloquium. A free shuttle will provide limited tranport
between SDSU and the Comfort Inn Saturday and Sunday.


Dr. Steven Barlow
EM Facility/Biology Dept.
San Diego State University
San Diego, CA 92182-4614
phone: (619)594-4523
fax:(619)594-5676
email:sbarlow-at-sunstroke.sdsu.edu






From: Ian MacLaren :      MACLARIZ-at-novell2.bham.ac.uk
Date: 7 Mar 1996 11:13:37
Subject: Re: TEM: preparation of ceramic fibres

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential
"Childress, Charles P. (Chip)" {cpc4-at-NIOBBS1.EM.CDC.GOV} ,
daughecc-at-UCBEH.SAN.UC.EDU, LLOYDPF-at-ml.wpafb.af.mil
Message-Id: {960306140912.626-at-cliff.ml.wpafb.af.mil.0}

Thanks to all who have responded speedily to my enquiry about preparation of
ceramic fibres for TEM. I will be trying some of these ideas in the next
few days.

Thanks again

Ian MacLaren,
IRC in Materials for High Telephone: 0121 414 3447
Performance Applications, FAX: 0121 414 3441
The University of Birmingham, email: I.MacLaren-at-bham.ac.uk
Birmingham B15 2TT,
England




From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Thu, 7 Mar 1996 09:17:20 GMT
Subject: Re: photographic processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Considering that photography is still so important to our profession, it is
} likely that in this group are a few critical individuals with personal
} experiencee using dry-to-dry black and white photographic processors. Would
} anyone care to recommend a counter-top processor that they are happy with? I
} wouldn't object to knowing which ones to stay away from either.
}
} Many thanks,
}
} Doug Keene
} Shriners Hospital for Crippled Children
} Portland, Oregon
}
} } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } We have
been using the Ilford 2150 RC for over a year, maybe two by now.
The results are quite good, however there have been mechanical problems with
it from the get go. We had to replace bearings early on and have had to
dismantle and re-assemble the rollers over and over to keep it from
squeaking and sqealing and grinding. So it has been a mixed bag for us. We
do about 10,000 prints a year and it has been a real time-saver too
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-





From: David.Rothbard-at-ipst.edu (David Rothbard)
Date: Thu, 7 Mar 1996 10:04:19 -0500
Subject: Publisher of S. Klosevych Series?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02130514ad64aa54fad2-at-[199.77.235.102]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Does anyone know who published, and the current source, of a series of
tutorial papers called "Microscopy and Photomicrography"? They were
written by Stanley Klosevych of the University of Ottawa.

Thank you in advance,
David Rothbard

--
Institute of Paper Science and Technology






From: kennedy-at-nsi.edu (grace kennedy)
Date: Thu, 7 Mar 1996 08:19:37 -0800
Subject: microscope disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: grace-at-popmail
Message-Id: {ad64bbaa000210040144-at-[198.147.151.19]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Fellow microscopists, what is the best and most efficient way to get rid of
an old TEM?? I see many for sale, etc., but have never seen replies. We
have an old Hitachi H-500 STEM, maintained by Hitachi until last year when
the $ ran dry. It's been used strictly as a biological TEM for the last 7
years or so. I would hardly thnk it would pay anyone to have it
disassembled and shipped, and it can't be left in place. Thanks in advance
for any input.






From: Elinor Solit :      cambrex-at-world.std.com
Date: Thu, 7 Mar 1996 12:14:11 -0500 (EST)
Subject: Re: MSA Meeting Info. Wanted

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lee,

The good people at the MSA office will be glad to help. Ask for Sharon or
Larry. Their number is 800-538-3672

See you at MSA.

Ellie Solit
The Microscope Book

On Wed, 6 Mar 1996, Lee Wagstaff wrote:

} Hi Gang,
}
} Anyone out there know where to obtain some comprehensive info on the MSA meeting coming
} up this August?? Any assistance would be appreciated. Thanks. Lee.
}




From: Linda Fox :      lfox1-at-wpo.it.luc.edu
Date: Thu, 07 Mar 1996 10:12:24 -0600
Subject: photographic processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603071725.LAA11394-at-Sparc5.Microscopy.Com}

The best processor out there is the Mohrpro. We have been using one
for about 8 years and recommend it when ever we can. It is easy to
use and maintain, 2 min. dry to dry, permanent (not stabilized) and
uses RC paper and Kodak Fixer. We do purchase the developer from
Mohr. The current cost for paper up to 8" width is $4295, the
Mohrpro8. and 14" width $4800, the Mohrpro14. (They do sell
rebuilt machines as well). They stand behind their product and I can
only say great things about Mohr.
For information contact:
Bob or Jim Jackson (tell them Linda at Loyola sent you!)
Mohr Enterprises
65 E. Palatine Rd. Suite 103
Prospect Heights, Il 60070
1-847-465-0048

Linda Fox Loyola University Medical School
lfox1-at-wpo.it.luc.edu





From: knoff-at-lipovx.lbl.gov (Laura Knoff)
Date: Thu, 7 Mar 1996 11:01:26 -0800
Subject: NCSM Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603071758.LAA11441-at-Sparc5.Microscopy.Com}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


WHAT: The Spring meeting of the Northern California Society for Microscopy.

WHEN: Thursday, March 14 from 5:30 to 9:30 p.m.

WHERE: Genentech, 460 Point San Bruno Blvd., So. San
Francisco, Ca

Speakers: Dr. Wen Lu Li, Genentech -

"Microscopy as a Tool in the Development of Theraputic Molecules"
&
Dr. Ming Pan, Gatan -

"High Resolution Transmission Electron Microscopy of Zeolites"

The Northern California Society for Microscopy (NCSM) welcomes new
members. Anyone interested in microscopy of any type, from atomic force to
optical will benefit from joining the NCSM. We meet four times a year to
have dinner and hear about new developments in microscopy. Dues are very
reasonable at only $15 a year for regular members, and $8 for students. If
you are interested in joining, or knowing more about NCSM, please contact
Kent McDonald by phone at 510-642-2085 during normal business hours, or by
email at: klm-at-uclink4.berkeley.edu If you are interested in attending the
Mar. 14 meeting described above, please call 510-486-4088 no later than
Mar. 8 so we can reserve the appropriate number of dinners. Complimentary
drinks will be provided by MicroStar Diamond Knives and RMC and the dinner
will be $15 for members, $7.50 for student members. Please join us.








From: William R Oliver :      oliver-at-ipas4.afip.mil
Date: Thu, 7 Mar 1996 12:31:19 -0500 (EST)
Subject: Re: microscope disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



What about museums and such? I know that there are technology/medicine
museums scattered about which might have a use. I know that the National
Museum of Health and Medicine here at the AFIP has a microscope
collection, and I believe the Smithsonian might as well. I assume that
there would be other museums in the country which would also benefit from
such donations.

billo

On Thu, 7 Mar 1996, grace kennedy wrote:

} Fellow microscopists, what is the best and most efficient way to get rid of
} an old TEM?? I see many for sale, etc., but have never seen replies. We
} have an old Hitachi H-500 STEM, maintained by Hitachi until last year when
} the $ ran dry. It's been used strictly as a biological TEM for the last 7
} years or so. I would hardly thnk it would pay anyone to have it
} disassembled and shipped, and it can't be left in place. Thanks in advance
} for any input.
}
}
}




From: LSFY69A-at-PRODIGY.COM (MR KARL H BERGER)
Date: Wed, 06 Mar 1996 17:15:40 EST
Subject: microscope for sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Zeiss EM-109 for sale, good condition, $ 5,000.- or best offer.
Call 908-370-8082





From: LSFY69A-at-PRODIGY.COM (MR KARL H BERGER)
Date: Wed, 06 Mar 1996 17:15:40 EST
Subject: microscope for sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Zeiss EM-109 for sale, good condition, $ 5,000.- or best offer.
Call 908-370-8082





From: MelanieOwl-at-aol.com
Date: Thu, 7 Mar 1996 17:23:33 -0500
Subject: EDS Detectors for FE-SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Is it necessary to get a 30 mm window for an EDS detector to be used on a
Schottky emitter FE-SEM in order to get enough counts, or is a 10 mm okay?

Thanks in advance,
Melanie Behrens
Texaco, Inc.
Fuels and Lubes Research Dept.
Beacon, NY
behrema -at- Texaco.com or MelanieOwl -at- aol.com




From: Deborah Holmberg :      dlholmberg-at-ucdavis.edu
Date: Thu, 7 Mar 1996 14:33:39 -0800 (PST)
Subject: Scanning 96, student volunteers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello All,
The Scanning 96 Meeting in Monterey, CA on 9-12 April is looking for
students to monitor the sessions, ie. run the slide projectors, adjust
lights and hand out information. The student will have the registration
fee waived, in appreciation of helping. For information, please contact
me, or the Scanning 96 office at FAMS, Inc., PO Box 832, Mahwah, NJ
07430-0832. E-mail: fams-at-holonet.net
Thank you,
Debe Holmberg
USDA-ARS
916.752.9021
916.752-4604 (fax)




From: VCRVINCE-at-aol.com
Date: Thu, 7 Mar 1996 20:39:56 -0500
Subject: Si Colors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dan,

Si color changes related to us by Tom Cass in '81 when VCR GROUP designed the
first DIMPLER, D100 were:

Garnet10-12um, Wheaty brown 7-9um, yellow 4-6um, white 1-3um, fringes {1um


The basic dimpler design came from his lab at Hewlett Packard Deer Creek
Labs, EMSA Bullentin 10(1) 1980, p.66, Addendum 10(2) 1980, p.93.

The colors depend on the intensity and type of lighting, Si dimpled and
viewers color perception. Specimen thickness can be measured quickly and
accurately on the DIMPLER specimen platen with a good optical microscope.
This measurement technique is especially helpful with specimens that are not
light transmitting.

V. L. Carlino
VCR GROUP, Inc.







From: Steven Schwarz :      sschwarz-at-morgan.ucs.mun.ca
Date: Thu, 7 Mar 1996 19:14:08 -0330 (NST)
Subject: LH-zeiss parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi-

I have a Zeiss Universal 'M' microscope which is missing a few
parts, which are now VERY expensive (from Zeiss).

I require a REFLECTOR for a 'Vertical Illuminator IIC'; either a
H-PL, or H-PR will do. Only the reflector is required as I have the
housing. Does anyone have one of these ? Or does anyone know if it is
possible to 'make' one using a 'double sided semi-transparent mirror' ?


Many thanks

steve


==============================================================================
Steve Schwarz sschwarz-at-morgan.ucs.mun.ca
Dept. of Earth Sciences
Memorial University of Newfoundland
Newfoundland
CANADA
A1B 3X5
1-709-737-8142
-737-2589 FAX
******************************************************************************




From: dianavd-at-eye.usyd.edu.au (Diana van Driel)
Date: Fri, 8 Mar 1996 10:47:20 +1100
Subject: Re: microscope disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: diana-at-pc-0.eye.usyd.edu.au
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Fellow microscopists, what is the best and most efficient way to get rid of
} an old TEM?? I see many for sale, etc., but have never seen replies. We
} have an old Hitachi H-500 STEM, maintained by Hitachi until last year when
} the $ ran dry. It's been used strictly as a biological TEM for the last 7
} years or so. I would hardly thnk it would pay anyone to have it
} disassembled and shipped, and it can't be left in place. Thanks in advance
} for any input.


Metal recyclers are an option. Taking parts home as unusual potplant
holders (bits of the column), paperweights (the lead glass), decorations
etc is fun. Perhaps a local school would like it.


Diana van Driel
Dept Ophthalmology C09
Sydney University 2006
NSW, AUSTRALIA






From: Stephen Edgar :      s.edgar-at-auckland.ac.nz
Date: Fri, 8 Mar 1996 10:54:49 NZS
Subject: Re: Publisher of S. Klosevych Series?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Does anyone know who published, and the current source, of a series
} tutorial papers called "Microscopy and Photomicrography"? They were
} written by Stanley Klosevych of the University of Ottawa.
}
} Thank you in advance,
} David Rothbard


Dear David,

I believe these tutorials have been published as a single volume, but
I do not know the details of that.

They were also published as a series of six papers in the Journal of
Biological Photography in 1974 and 1975. The details are:

Part I vol. 42:part 3, pp 123-131
Part II 42:4, 147-160
Part III 43:1, 30-38
Part IV 43:2, 51-69
Part V 43:3, 119-139
Conclusion 43:4, 187-188


About 24 degrees Celsius and not a cloud in the sky. Who needs Hawaii?


Regards

Stephen Edgar

Electron Microscope Unit, Pathology Department
School of Medicine
University of Auckland
Private Bag 92019
Auckland
New Zealand

email address: s.edgar-at-auckland.ac.nz
Phone : +64-9-3737599 extn 6473 (GMT + 12h)
Fax : +64-9-3737459




From: Roar Kilaas :      Roar_Kilaas-at-macmail2.lbl.gov
Date: 7 Mar 1996 13:44:34 U
Subject: Workshop on Quantitative HR

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1385903893.39621-at-macmail2.lbl.gov}

REGARDING Workshop on Quantitative HRTEM

Announcement of Workshop/Symposium
-------------------------------------------

The National Center for Electron Microscopy at the Ernest Orlando Lawrence
National Laboratory will be hosting a 3 day workshop on Quantitative High
Resolution
Transmission Electron Microscopy, April 18 - April 20, 1996.
This workshop is designed to address many important issues relating to
obtaining
quantitative information from experimental data in HRTEM. The workshop will
begin
with discussing the many steps required in order to ensure meaningful
quantitative
information, starting with the recording of the experimental data,
transforming
recorded information into digital form, correcting for anomalies of the
recording media and processing of the data. Procedures for obtaining
quantitative measurements
of important parameters such as chemical composition, grain boundary
expansion/contraction, strain and lattice parameters will be outlined,
including a
discussion of the accuracy of the measurements.
The workshop will cover problems associated with matching of calculated images

to experimental data and address the possibility of creating standards for
reporting
goodness of fit between experimental and simulated images and confidence
levels for
refined structural models. There will be one day devoted to reconstruction
methods
in HRTEM including a discussion on the limits of current reconstruction
methods.

Registration for the meeting is $150. Total number of participants is limited
to 40.
Application for registration can be obtained by contacting Gretchen Hermes
(see
contact information below) or by mailing or faxing (to Gretchen Hermes) the
application
form obtainable from the web-site listed below.

Gretchen Hermes
tel: 510-486-5006
fax: 510-485-5888
email: GHermes-at-lbl.gov

Total number of participants is limited to 40.

For information on registration and further details on the workshop, please
see the
web site:
http://ncem.lbl.gov/NCEM/workshops.html
--------------------------------------------------
Roar Kilaas
NCEM/MSD, MS 72-207
Lawrence Berkeley National Laboratory
1 Cyclotron Road,
Berkeley, CA 94720
tel: 510-486-4618
fax: 510-486-5888
email: r_kilaas-at-lbl.gov






From: mager-at-unixg.ubc.ca (Mary Mager)
Date: Thu, 7 Mar 1996 21:14:44 -0800
Subject: Re: Publisher of S. Klosevych Series?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: mager-at-pop.unixg.ubc.ca
Message-Id: {v01510100ad656ffddb2f-at-[137.82.220.103]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Does anyone know who published, and the current source, of a series of
} tutorial papers called "Microscopy and Photomicrography"? They were
} written by Stanley Klosevych of the University of Ottawa.
Dear David,
The Stanley Klosevych series was published in the Microscopical Society of
Canada Bulletin. The Bulletin Editor, Caroline Emerson (email:
cemerson-at-kean.ucs.mun.ca), would be the best place to try for copies. I
don't remember exactly when the series was published. There is also a very
good book by him: "Principles and Practice of Microscopy and Scientific
Photography", which can be ordered on the MSC home
page:http://gause.biology.ualberta.ca/craig/hp/MSC.SMC.Mic.Soc.hp. or you
can link from the MSA/Microscopy site.
Luck,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Eng, UBC
309 - 6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648, fax: 604-822-3619
email: mager-at-unixg.ubc.ca






From: Margaret H Malay :      malay-at-csd.uwm.edu
Date: Thu, 7 Mar 1996 23:44:48 -0600 (CST)
Subject: Midwest Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopy of Surfaces and Interfaces
Midwest Microscopy and Microanalysis Society Workshop
UWM Microscopy Open House

April 26, 1996

Please register by April 15, 1996
More information and online registration at
http://www.uwm.edu/Dept/LSS/meeting.html

PROGRAM
Golda Meir Library Conference Center
University of Wisconsin - Milwaukee

Opening Ceremonies
8:45 - 9:00

Morning Session

9:00 - 9:30
J. Mansfield
University of Michigan - Ann Arbor
"Materials Science and Biological Applications of Environmental Scanning
Electron Microscopy"

9:30-10:00
B. Tonner
University of Wisconsin - Milwaukee
"X-ray Microscopy for Chemical Analysis of Surfaces"

10:10 - 10:15 Coffee Break and Mounting of Posters
Sponsored by College of Letters and Sciences

10:15 - 10:45
T. Kelly
University of Wisconsin - Madison
"Three-Dimensional Atom Probe Microscopy"

10:45-11:15
J. Nogami
University of Wisconsin - Milwaukee
STM: "Studies of Metal Growth on Silicon Surfaces"


11:15-11:45
L. D. Marks
Northwestern University - Evanstone
Surface HREM: "Old techniques on Old Surfaces - Surprising New Results"

Open House Lunch

11:45 - 1:00
Lunch sponsored by Nissei Sangyo America and Hitachi Scientific Instruments.
Poster Session

1:00 - 1:45
Poster viewing and poster competition
Poster prizes are sponsored by the MMMS.

Afternoon Session

1:45-2:15
J. M. Gibson
University of Illinois - Urbana
TEM: "All you ever wanted to know about building your own expensive and
complicated microscope, and then waiting five years until it finally works"

2:15 - 2:45
S. Babcock
University of Wisconsin - Madison
TEM/HREM: "Electron Microscopy Studies of Grain Boundaries in Bicrystals"

2:45 - 3:15
V. Dravid
Northwestern University - Evanstone
"Analytical Electron Microscopy and Holography of Interfaces: Making a
Mountain out of a Mole Hill"

3:15 - 3:30 Coffee Break and Judging of Posters
Sponsored by Laboratory for Surface Studies

3:30 - 4:00
N. Browning
University of Illinois at Chicago
"High Angle Annular Dark Field in a Dedicated STEM"

4:00 - 4:30
K.L. Merkle
Argonne National Laboratory
"Relaxation Modes in Metal and Oxide Grain Boundaries"

4:30- 4:45
Poster Awards Ceremony and Closing Address
M. Gajdardziska-Josifovska, Workshop Organizer

Open House Tour

REGISTRATION FEES:
Registration is free for the members of the Midwest Microscopy and
Microanalysis Society. The workshop registration fees for non-members are
same as the membership fees and can be paid at the conference site (by check
or cash):
Regular Member ($10)
Student Member ($5)

POSTERS
All lectures are by invitation only. Poster contributions by students,
researchers and microscope manufacturers are welcome. There are no size
restrictions for the posters, however, they should be mounted on a sturdy
backing suitable for display on easels. The poster must be relevant for the
broad topic of microscopy of surfaces and interfaces.

POSTER COMPETITION

Graduate students are particularly encouraged to present their work and
they are eligible to enter a poster competition. One grand prize and two
first prizes will be given by the Midwest Microscopy and Microanalysis
Society. The grand prize must be used by the winner to attend a microscopy
related session at a major national/international conference.

Grand Prize: $500
1st Biological Sciences:$100
1st Physical Sciences: $100








From: Tony Bruton :      bruton-at-emu.unp.ac.za
Date: Fri, 08 Mar 1996 11:01:54 +0200
Subject: Re: microscope disposal -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s14013be.065-at-smtp.unp.ac.za}
X-Mailer: Novell GroupWise 4.1

We have just disposed of an old microscope - a HITACHI HU-11E.

After advertising it on the Net and finding no takers we dismantled it. We
kept things like DP and RP pumps, transformers, motors and guages but
found the recovery of scrap metal very limited. The chassis fetched the
princely sum of R33.00 from the local scrap dealer. Perhaps as a
teaching facility the greatest value was the fact that we now have a
wonderful assortment of dissected lenses, aperture holders, electron
gun etc to use as teaching material.

I don't believe that it is a kindness to donate a 27 year old fragile
valve-powered instrument such as this to a school or suchlike, the
maintenance responsibilities outside of expensive, specialised hands
are overwhelming.

An enquiry to the Japanese suppliers gave us a one-liner which we
have already framed - "Please throw away TEM as scrap metals"

Sobering moment seeing the frame of our beloved ex-TEM swinging
from a crane in the scrapyard before being dumped onto a pile of old oil
drums........

Tony Bruton
Centre for Electron Microscopy
University of Natal
Pietermaritzburg
KwaZulu Natal
South Africa





From: Wharton Sinkler :      sinkler-at-dvibm3.gkss.de
Date: Fri, 8 Mar 1996 11:32:46 +0000 (MET)
Subject: alignment of CM30

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello all:

I am using a Philips CM30ST for high resolution imaging. I have had a
problem performing the critical objective lens direct alignment "current
rotation center". Ideally, one should obtain a stationary point on the
specimen, rotating about its center at high magnification with a
low-frequency wobble applied to the objective lens current. I can get
the stationary point on the specimen reasonable satisfactorily.
The probjem is that the beam moves fairly significantly when the objective
lens wobbler is on.

I have performed the coma free direct alignment, which should ensure that
the beam is travelling along the optic axis of the objective lens. This
works well. However, it seems to me that if wobbling the objective lens
current can shift the beam, it can't be travelling the optic axis.

Is there any relatively quick way to correct the problem of beam shift
with objective lens current change? I can imagine that a complete
realignment of the microscope might get rid of the problem. However,
this is both very time-consuming and also not very instructive as to the
cause of the problem.

Wharton Sinkler

==================================================

Dr. Wharton Sinkler
GKSS Forschungszentrum Geesthacht
Abteilung WA
D-21502 Geesthacht
Germany

tel: (49)(4152)872542
fax: (49)(4152)872534




From: Ian MacLaren :      MACLARIZ-at-novell2.bham.ac.uk
Date: 8 Mar 1996 10:35:25
Subject: Re: TEM: preparation of ceramic fibres

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: Microscopy-at-Sparc5.Microscopy.Com

Dear all,
I have received a whole bunch of suggestions about the above subject and
someone has asked me if I would share a summary of the responses with the
whole list, this is given below. Thanks again to all who responded. We
have decided to embed the fibres in epoxy as suggested by some of the
contributors, dimple and ion mill (at a low angle). I don't think that we
have a tripod polisher but I will be checking that out.

____________________________________________________________________________
My original message:

Dear all,
Has anyone out there prepared non-conducting ceramic fibres for TEM, either
across or along axis? If so, how did you do it?
We have some mullite fibres we would like to look at and although we have
some ideas about how to prepare specimens we would appreciate any ideas or
feedback on this topic.
____________________________________________________________________________
We have been routinely preparing ceramics fiber tows (Nextel, Nicalon, HPZ,
etc.) with and without ceramic coatings for TEM characterization for about
two years now. After much trial and error with a number of methods (some in
the literature) we have settled on a rather efficient method of producing
high-quality sections. We are just finishing a short communication on this
technique, which I could send to you when completed.

First, the tows are vacuum-impregnated with a high-temperature epoxy, such
as G-1. It is important that the fiber-volume fraction be high, which can
be achieved by either pulling many tows through the tip of a pipette, or
alternatively using shrink tubing. The epoxy is then allowed to cure.

The key to obtaining a good thin section with a large electron transparent
area is to minimize ion-milling time, which requires mechanically thinning
the sample to thicknesses less than 5 um, ideally less than 1 um. We use a
tripod polisher with 3M diamond lapping films to minimize differential
polishing and surface relief. The epoxy-impregnated tows have been thinned
both axially and transversely; either way wil produce good specimens.
Again, you want to thin the specimen as much as possible prior to ion
milling; if the specimen is too thick the epoxy will be milled away before
the fibers and any coating becomes electron transparent, with the fibers
usually falling out. We also need the epoxy to bound the coating for
coating thickness determination.

The specimen is then mounted on a Cu washer, and maybe a grid, and ion
milled at low angles for 30-60 min.

For more details please contact me off line.

Dr. Michael Cinibulk cinibumk-at-ml.wpafb.af.mil
Wright Laboratory/MLLM
Wright-Patterson Air Force Base, Ohio
____________________________________________________________________________
Do you have a tripod polisher in your prep lab.

We have a way of prepping fibers using a tripod that worked nicely
for cross and longitudinal sections of SiC fibers.

Ron Anderson
____________________________________________________________________________
I assume that you are aware of the difficulties in doing this. I know of
two methods which work very well and are based on other, relatively
well-known techniques for producing cross-sections.

A) embed in resin and ultramicrotome cross-sections; this is only
convenient if you have someone who does ultramicrotomy already, in which
case it is probably the simplest.

B) sandwich with epoxy between two silicon wafers and then cross-section
'normally,' just like a VLSI device; by visually monitoring optical
transmission through the silicon, you can thin the composite structure well
below 10 micrometers before ion-milling (which in turn helps mitigate
differential milling effects). This works for almost everything. The same
basic technique could be used to thin the fiber along the axis.

Please let me know if something easier comes along, since I have to be
doing this as well...

Daniel L. Callahan
Assistant Professor of Materials Science
Rice University
http://www.owlnet.rice.edu:80/~dlc/
____________________________________________________________________________
I had pretty good luck preparing ceramic fibers (mullite,
SiC, Al2O3/ZrO, and others), of 10 to 30 micron diameter,
for TEM analysis. The method is very simple, and consists
of first making a composite out of the fibers and an epoxy,
then thining down as with a bulk sample (polishing,
dimpling, ion milling). Things to keep in mind:

1. If the fibers are short and tangled as a small cotton
ball, saturate the epoxy with the fibers mixing very well.
The fibers will brake but that doesn't matter since is
unlikely that you will be interested in crystalline defects.
Squeeze the composite between two teflon or other
non-sticking sheets using glass slides for backings.


If the fibers are long (more than 3mm) you can lay down
a thin layer of epoxy on a teflon sheet, and the fibers on
top, one at a time or in bundles, in a single direction to
cover a strip 3mm wide x length of the fibers. Then squeeze
between two glass slides and another teflon sheet covered
with epoxy. Cure the epoxy as per manufacturer instructions.
In both cases you will be able to form a very thin composite
that requires a little grinding and polishing on both
sides before punching out 3mm discs, which can be further
polish to 100 micron. Dimple as usual (with diamond paste)

2. Epoxy: try with whatever you have available. Some
people prefer conductive epoxies. Initially, I used several
conductive and non conductive products from TRA-CON
(Medford, MA, 508-391-5550), but for some time now I have
settled on G1 epoxy from GATAN (412-776-5260).

3. Ion Milling: the mullite I worked with milled very fast
probably because it was very porous. Other ceramics mill
very slowly compared to the epoxy. This is a problem that
can be alleviated using a LN2 stage in the ion mill, to slow
down the milling rate of the epoxy, and having as much
fiber content as possible in the mixture.

4. Even if you use a conductive epoxy you may have to
carbon coat the thinned samples.

Good Luck!

Augusto Morrone
Univ. of Florida
Materials Science and Engineering
P.O.Box 116400
(352) 392-6985
amorr-at-mse.ufl.edu
____________________________________________________________________________
been using the Tripod polisher and a unique technique for tightly binding
fibers together. They have been working with SiC fibers, but I think that it
would work for you. Their Email addresses are
CINIBUMK-at-ml.wpafb.af.mil and SCHELTFJ-at-ml.wpafb.af.mil

- -Scott Walck
____________________________________________________________________________
I think that someone here used a Gatan PIMS to do this several years ago
with some success. The length of the fiber went across the 3mm washer.
John Hunt
____________________________________________________________________________
Over the years I've prepared various fibre materials in various ways. These
all should give results; the quality varies, but then so does the effort
required.

1) If the fibres are thin to begin with, pop them onto a support film
covered
grid, carbon coat, and voila!

2) If you have experience with ultramicrotomy, embed and section small
areas.

You may require a coupling agent to promote resin adhesion, and the result
of sectioning will be a mass of shattered fragments (unless the fibres are
amorphous), but then you might also get a huge amount of viewable area free
of chemical contamination.

3) Mix the fibres into a particularly hard resin such as Petropoxy which you
can then prepare (cut, grind, dimple, ion mill or tripod) as a monolithic
lump. I'm not aware of any resin hard enough to thin at the same rate as a
ceramic, but I've prepared SiC fibres this way with success.

4) Create a composite by electroless deposition of metal (usually Ni) which
you can then prepare in the ion mill. This one gives by far the best result
in my experience, but is also the trickiest and most potentially time
consumming. A short electroless plate to make the fibres conductive ,
followed by electrolytic deposition in standard plating baths is a good
option if you have access to such. The method I've used was adapted from
metal powder prep. techniques.

Let me know if you'd like further information on any of this.

###############################################################
# #
# Don Steele STEELE-at-KRDC.INT.ALCAN.CA #
# ALCAN INTERNATIONAL #
# Kingston Research and Development Center #
# P. O. Box 8400 #
# Kingston, Ontario Canada K7L 5L9 #
# #
###############################################################
____________________________________________________________________________
The answer is simple (at least in principle!) - embed them in a hard
epoxy resin or acrylic, then section them in an ultramicrotome,
collect on fine-meshed grids, pop in the TEM, and ---presto, nice
uniformly thin fiber cross-sections held together by the resin.
There will undoubtedly be some tearing of the section and/or
individual cross-sections. The fibers will be randomly dispersed so
that you'll get many orientations (inlcuding some near-longitudinal
slices, if they're useful). Be careful of beam effects - most
embedding materials are somewhat to terribly beam sensitive.

Sound too easy to be true? You're right, ultramicrotomy is somewhat
of an art, but can accomplish wonders in materials science. A good
reference is an overview by yours truly:

T.F. Malis and D. Steele, Specimen Preparation for Transmission Electron
Microscopy of Materials, MRS vol 199, Materials Research Society (1990) p.3.

If you don't have any MRS Proceedings there, I have a few copies left and
can send one, but only if you're seriously interested, as they're in short
supply. The people at UMIST in the Corrosion Center are experts in
this area as well. I tend to act as a information bank for this technique,
so let me know how you fare - I'm always looking for new references to
quote.

Tom Malis
Group Leader - Materials Characterization
Materials Technology Laboratory
Natural Resources Canada (Govt. of Canada)
Ottawa, Ontario
ph.: 613-992-2310
FAX: 613-992-2310
e-mail: tom.malis-at-cc2smtp.emr.ca
____________________________________________________________________________
This may or may not be applicable, but with the high-voltage EM
(lower interaction cross-section) and very low beam currents, we have been
able to image non- or poorly-conducting specimens without charging or heat-
ing problems. If this holds true for your instrument (e.g. you have an
IVEM), the advantage is that there may be no preparation required; the
disadvantage is that with these low doses you will either need a very sen-
sitive recording medium--LoDose or other x-ray film, image plates or in-
tensified or slow-scan CCD--or long exposures. The large grain or speci-
men drift could make the images useless. Good luck.
Yours,
Bill Tivol
____________________________________________________________________________
I am a little late in responding to your inquiry on ceramic fibers; My
apologies. Since some of the other responses mentioned using the wedge
technique for preparing thin TEM sections, I wanted to mention that BUEHLER
KKB, with whom I am affiliated, is a supplier of a wedge polishing tool such
as those mentioned. I previously was Applications Engineer for South Bay
Technology who offered the first commercially available tripod based
polisher. However, I have since come to work for BUEHLER, LTD in the US,
and we have made some changes to the IBM design in order to enhance the ease
of producing wedge samples. We also offer a complete line of diamond
lapping films for this polishing system.

If you would be interested in more information regarding BUEHLER's
MICROPRECISE(TM) Tripoint Polisher, I would be happy to have literature sent
to you, and/or have a salesperson contact you. If you would like more
information, please feel free to contact me directly by phone, fax or
e-mail.

Best regards,
Scott D. Holt
BUEHLER, LTD.
41 Waukegan Rd.
Lake Bluff, IL 60044 USA
Phone: (847)295-4546
Fax: (847)295-7942
102467.2752-at-compuserve.com
____________________________________________________________________________


Ian MacLaren,
IRC in Materials for High Telephone: 0121 414 3447
Performance Applications, FAX: 0121 414 3441
The University of Birmingham, email: I.MacLaren-at-bham.ac.uk
Birmingham B15 2TT,
England




From: David.Rothbard-at-ipst.edu (David Rothbard)
Date: Fri, 8 Mar 1996 08:03:36 -0500
Subject: S. Klosevych Series. Thank you.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02130501ad65e02aade5-at-[199.77.235.102]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Eight is enough. I'd like to thank everyone who responded with the
reference information on this series. You can stop responding now.

David Rothbard

--
Institute of Paper Science and Technology






From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Fri, 8 Mar 1996 06:41:37 -0600
Subject: Re: alignment of CM30

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Coma-free alignment (stationary image of amorphous carbon when applying
plus/minus tilts at high magnification, 500-700K) and current centering
(specimen stationary when oscillating the lens current) are completely
different. The reason is that the top and bottom parts of the objective
lens are never completely lined up, the mechanical tolerances are too
severe. Coma-free alignment finds the optic axis in a true sense
from the sample downwards taking into account the misalignment of the
two parts of the objective lens. The beam is still likely to shift
when the objective lens current is oscillated since this is largely
determined by the top part of the objective lens. As a test, you
can measure the difference in tilt between:
a) Coma-free alignment
b) Current-center alignment
c) HV-center (oscillating the voltage, equivalent to testing the
alignment of all the post specimen lenses)
d) The tilt such that the beam does not translate when the objective
lens current is oscillated.

By comparing d) and a) you can get a measure of the misalignment between
the top and bottom parts of the polepiece; comparing a), b) and c) you can
check if the other post specimen lenses are well aligned. If the difference
between d) and a) is large then you have a bad lens, and you will have
problems getting small probes for microanalysis without artifacts (e.g.
tails).




From: randy nessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Fri, 8 Mar 1996 10:11:28 -0600 (CST)
Subject: Re: EDS Detectors for FE-SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Thu, 7 Mar 1996 MelanieOwl-at-aol.com wrote:

} Is it necessary to get a 30 mm window for an EDS detector to be used on a
} Schottky emitter FE-SEM in order to get enough counts, or is a 10 mm okay?
}
} Thanks in advance,
} Melanie Behrens
} Texaco, Inc.
} Fuels and Lubes Research Dept.
} Beacon, NY
} behrema -at- Texaco.com or MelanieOwl -at- aol.com

Our 10mm does fine, as long as the apts and lens are set properly.

Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu





From: bsgphy3-at-uconnvm.uconn.edu (JIM ROMANOW)
Date: Fri, 8 Mar 1996 11:29:49 -0500
Subject: Re: Microscope disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510101ad660a6afa36-at-[137.99.40.87]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} an old TEM?? I see many for sale, etc., but have never seen replies. We
} have an old Hitachi H-500 STEM, maintained by Hitachi until last year when
} the $ ran dry. It's been used strictly as a biological TEM for the last 7
} years or so. I would hardly thnk it would pay anyone to have it
} disassembled and shipped, and it can't be left in place. Thanks in advance
} for any input.

Do you have an art department? Some sculpture students are inspired by old
machine parts and they usually handle the shipping quickly. About 3 trips
for a Volkswagon bus (They were rated as half ton vehicles).

Good Luck
Jim

Jim Romanow
Electron Microscopy Facility
Physiology and Neurobiogy Department
The University Of Connecticut
Storrs
bsgphy3-at-uconnvm.uconn.edu






From: gins-at-acs.bu.edu
Date: Fri, 8 Mar 1996 14:00:47 -0500
Subject: recycling TEM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential


On the subject, a friend uses the old vacuum pump from a Phillips for the
vacuum oven he uses for embedding in parafin and epoxy. I don't know
if it is qualitatively better thant a typical house vacuum line, but our
house vacuum has been known to be unreliable.

C.Ginsburg
Boston Univ. Med. School




From: GANTZ-at-med-biophd.bu.edu
Date: Fri, 08 Mar 1996 13:42:25 -0500 (EST)
Subject: Bench-top Photographic Processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Doug:
We've had our Mohr Pro8 for about 9 months and love it. This model came
highly recommended by other users in the Boston area who had used them
much longer than us. We now wonder how we put up with the tray method
for so long! If you want more info, E-mail or call me.
Don Gantz
Boston Univ School of Medicine
gantz-at-med-biophd.bu.edu
617-638-4017

PS. AGFA also makes one but have no details.




From: Joe D Geller :      geller-at-world.std.com
Date: Fri, 8 Mar 1996 16:06:29 -0500 (EST)
Subject: Objective lens for Leitz Ergolux

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are in need of a low power, 4X, objective for the Leitz Ergolux
optical microscope. I believe this is an infinity corrected optical
system. Might one be available?

Joe Geller
Geller MicroAnalytical Laboratory
426e Boston St.
Topsfield, Ma 01983-1212
508 887-7000 fax:887-6671




From: epicier-at-univ-lyon1.fr (Thierry Epicier)
Date: Mon, 11 Mar 1996 17:51:11 +0100
Subject: SIMPLY-S for HREM simulations on PCs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603081842.MAA20224-at-Spacestar.COM}

The set of softwares SIMPLY-S, including for the multislice-based programs
SHRLI, and dedicated to various TEM/HRTEM calculations/simulations on
PCs (in the field of Materials Sciences) has recently been updated.
SIMPLY-S is available from ftp.univ-lyon1.fr (PUB/DOS/HRTEM : README.TXT
and SIMPLY1-2-3.ZIP files ; SIMPLY3.ZIP contains the updated version, while
SIMPLY1-2.ZIP files do correspond to the previous version, to be abandoned
soon).

Any microscopist interested can download the freeware version. remind also that
SIMPLY-S is permanently updated (check the date in the readme.txt file).

Comments and remarks are welcome !


______________________________________________

Thierry EPICIER
GEMPPM-502, INSA de Lyon,
69621 VILLEURBANNE, France
tel : (33) 72 43 84 94 (83 85)
FAX : (33) 72 43 85 28
Email : EPICIER-at-CISMSUN.UNIV-LYON1.FR





From: mallamaci-at-goodyear.com (Michael Mallamaci)
Date: Mon, 11 Mar 1996 12:55:03 -0500
Subject: Missed the ISO Guide 25 Summary

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: t901378-at-rds163
Message-Id: {v01530505ad6a18a9fbe6-at-[163.243.12.63]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Folks,

I seem to have missed the summary of the ISO Guide 25 Experience thread.
Was it ever out there? If somebody could forward it to me I would
appreciate the effort.

Thanks in advance,

Mike

Michael P. Mallamaci
The Goodyear Tire & Rubber Company
Analytical Sciences D/415A
142 Goodyear Blvd.
Akron, OH 44305 USA
--------------------------------
mallamaci-at-goodyear.com
(330) 796-7436






From: hawkey-at-neuro.duke.edu (Larry Hawkey)
Date: Mon, 11 Mar 1996 14:48:10 -0500
Subject: Has anyone processed E-coli for TEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510109ad6a319f88e1-at-[152.3.72.63]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Some one wants to look at E-Coli in the EM. I have never worked with
bacteria before. Where could I find some good references on how to fix and
process them. Is there a standard fix or pH or anything I should be aware
of before I dive in.

any guidance would be most helpful.

Larry Hawkey
hawkey-at-neuro.duke.edu






From: dshubito-at-d.imap.itd.umich.edu (Dennis Shubitowski)
Date: Mon, 11 Mar 1996 14:05:50 -0500
Subject: Re: ZIP100 MAC/PC disk exchange easy if...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02120d00ad6a27b3c6e3-at-[141.211.157.94]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


(snip)

} We have not seen a utility to mount Mac formatted disks on DOS systems.
}
} Iomapple
} Iomega Technical Support

(snip)

I hold in my hand Conversions Plus from DataViz that converts files between
Mac and PC programs, including floppy disks, removable cartridges (SCSI) and
CR ROMS. I personally do not use it, but the good doctor in the office next
door uses it quite regularly as he runs a Windows NT server for both Macs
and PCs. The programs runs about $100 (US), but its smaller cousin MacOpener
allows you to use Mac floppies, removable media, and CD ROMS, but not to
convert. This runs about $50 (US).

The above programs apparently have done all they claim to do without a hitch,
whereas he had had problems with Mac-In-Dos, a program that claims similiar
capabilities.

Dennis Shubitowski
University of Michigan
School of Dentistry

(No connection to DataViz. I'm a Mac person!)






From: albrite-at-leonardo.net (Larry Albright)
Date: Sun, 10 Mar 1996 14:31:53 -0800
Subject: Members and Speakers wanted.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html









Los Angeles Microscopical Society (LAMS)

LAMS has an ongoing need for interesting speakers for our monthly meetings.=
Topics could relate to the technique, history, applications, subjects or =
art involving microscopes. We are open to anything that is interesting,=
even if remotely related. This could be a good way to practice a new=
speech or presentation before a friendly and interested group.
LAMS is a non-profit organization founded in 1938, dedicated to the study=
and practice of the science, art, and history of microscopy and related=
topics. Membership covers individuals with an active interest and=
experience in optical and electron microscopy and associated science and=
technology. LAMS has a membership of around 100 people, and an average=
attendance of 40. The club has a monthly newsletter with a column by the=
President Gil Melle, usually on different aspects in the history of the=
microscope with artful illustrations of classic designs. We have=
corresponding members all around the globe. The membership fee is $25.00=
per year which includes the monthly bulletin. To join call Dave Hirsch at 3=
10-397-8357.


At this time, we conduct our meetings at the Edison Complex 1721 22nd Street=
in Santa Monica California. The meetings are held every third Wednesday=
of the month. Members gather about 7:00 p.m. and the program starting at=
8:00 p.m. The program usually lasts an hour and a half. There is a short=
refreshment break after the program and the second activity ending at 11:00=
.

The meeting room has a wall screen, blackboard,Carousel projector, large=
projection video, and other items available. Please let us know your needs.

Please e-mail to the Program Chair, Larry Albright {albrite-at-leonardo.net} =
or call if you are able to speak to our group, or if you would like to=
simply attend a meeting. The meetings are open to anyone.=20


Larry Albright
Program Chair
(310-399-0865)
albrite-at-leonardo.net=20






From: BGIAMMARA-at-magnum.mco.edu
Date: Mon, 11 Mar 1996 16:57:01 -0500 (EST)
Subject: MT1 Parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi you-all. I have come across a OLD Porter-Blum MT1 Ultramicrotome,
but are missing collets, knife holders, & specimen trimming block. Anybody
out there have any hearts to donate spare parts to make this thing sing?
A copy of the operating instructions book would be helpful, too.
Thanks a lot,

Beverly Giammara




From: James Lausier :      yojimbo-at-itsa.ucsf.edu
Date: Mon, 11 Mar 1996 15:03:47 -0800 (PST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Please unsubscribe. Thank You




From: Ginger Baker :      lizard-at-okway.okstate.edu
Date: Fri, 8 Mar 1996 14:36:32 -0600
Subject: The Oklahoma Microscopy Society (OMS) Workshop in April

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0

On Friday, April 12, 1996 the annual Spring Workshop of the Oklahoma
Microscopy Society (OMS) will be held in Room 246 in the Noble
Research Center at Oklahoma State University (OSU), Stillwater,
Oklahoma.



The main focus of the workshop is "Digital Imaging and Image
Processing" and is being presented by Dr. David Bright, 1996 MSA
Speaker and a Research Chemist from the National Institute of
Standards and Technology in Gaithersburg, MD. The workshop will
include some of the following topics:

Image Processing for Microscopists: Practice and Pitfalls Image
Processing for the Macintosh: (free software available)
What are digital images and why would one want to process them?
Introduction to NIH Image and MacLispix
Introduction to Image Processing
-aspect ratio- "correction" for tilt or non square pixels -background
subtraction and normalization
-direction of gradient
-euclidian distance map
-mathematical morphology
-linear hough transform



Other items of business:

(1) Student Best Electron Micrograph Contest: Don't forget to
bring those prize-winning pictures with you to the workshop. Electron
micrographs will be judged by attendees at the workshop. The winner
will receive a $50.00 award and the micrograph will appear on the
cover of the Fall 1996 OMS Newsletter. Second place winner will
receive an award of $20.00.

(2) The Fall 1996 Technical Meeting in conjunction with the
Oklahoma Academy of Science will be discussed. This year it will be
held at the Oklahoma State University-OKC on Friday, November 8, 1996.
Abstracts are due Monday, September 16, 1996.

(3) Final call for the submission of micrographs suitable for the
new OMS t-shirt. Bring your entries to the spring workshop or mail
them to Ginger Baker, Sec. before that date. This contest is not
limited to students. The winner will receive a first edition of this
T-shirt absolutely free. Remember that the T-shirts should be
considered part of OMS' outreach efforts: they will be seen mostly by
the non-microscopist public. In the now-famous words of our OMS
president, "No hexamethylchickenwire lattice structures, please!".

(4) Discussion on the OMS web page expertly compiled and edited by
Dr. Scott Russell, University of Oklahoma.

Address: http://www.uoknor.edu/electron/oms/

From this site, you can travel to a variety of pages for up-to-date
microscopy news and information!

(5) Various meetings and shortcourses coming up during the year.
Also, updates on the gradeschool microscopy kits and their use in the
elementary classrooms.



Preregistration for the OMS Spring Workshop is $5.00 for members,
$10.00 for student non-members, and $15.00 for non-members. After
April 10 add a $5.00 late fee. On site registration begins at 8:30
a.m. Lunch is included in the fee.



Lunch: smoked turkey and brisket with barbecue sauce, cole slaw, baked
beans, relish, wheat/white bread, and peach cobbler.



Parking: Lot 96: Each person driving to the workshop must have a
visitor parking permit to hang in the vehicle with the designated lot
number.

For more information contact either:

Wanda L. Edwards Ginger R. Baker
President-Elect, 95-96 Secretary/Treasurer, 94-96
Email: kmk285-at-vms.ucc.okstate.edu Email: lizard-at-okway.okstate.edu
Phone: (405) 744-7415 Phone: (405) 744-6765
FAX: (405) 744-5275 FAX: (405) 744-8263





From: Larry D. Ackerman :      mishot-at-itsa.ucsf.edu
Date: Mon, 11 Mar 1996 16:09:37 -0800
Subject: photographic processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {1.5.4b11.32.19960312000937.006680ac-at-itsa.ucsf.edu}
X-Sender: mishot-at-itsa.ucsf.edu
X-Mailer: Windows Eudora Light Version 1.5.4b11 (32)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I have used an Agfa system, chemistry and paper for 12 years. The processor
was not sufficiently robust nor reliable for us. In any case it is no longer
available. The Agfa Variable Contrast Premium paper is my favorite emulsion
for scientific work. I recommend that you consider the Durst Printo system.
It is modular so you can build it to suit your own needs. It is my choice of
currently available systems.





From: Hand-at-nso1.uchc.edu (Hand,Arthur)
Date: Mon, 11 Mar 1996 09:29:05 -0500
Subject: TEM and SEM - Facility User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would appreciate any information on current user fees at university EM
facilities. My institution has requested a comparison of our fees with those
of other facilities. Thanks for your help.

Arthur R. Hand, D.D.S.
Director, Central EM Facility
Univ. of Connecticut Health Center
MC 1610
Farmington, CT 06030
Tel: 860-679-4085
Fax: 860-679-4078
e-mail:
hand-at-nos1.uchc.edu





From: Peters-at-BSAC.UCHC.EDU (Klaus-Ruediger Peters)
Date: Mon, 11 Mar 1996 08:46:01 -0500
Subject: ZIP100 MAC/PC disk exchange easy if...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d00ad63859a9ec9-at-[155.37.2.10]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi, I found some useful information on ZIP DRIVES FOR MIXED ENVIRONMENT.

The recent discussion on ZIP installation problems were not very
encouraging but most problems may have had their cause in inadequate
drivers and software.

Iomega's ZIP drives (not involved in any way with the company) are very
useful for image transfer between Mac and PC, especially 12-16 bit images
which do not fit on a floppy, session files, scanned images, image
processing derivatives from image files etc., as well as image exchange if
FTP is not available.

The following technical details come from AOL/Iomega Forum/Message
Board/Topic: ZIP Technical for PC/Mac. Also, all newest Iomega drivers for
Mac, DOS, Win can be copied from this site. I hope it helps. Best regards
Klaus.

********************************************************************************

} Is there any way for my Mac/Zip disks to be accessed on a PC /Zip drive or
} visa versa?

}
}
} Bernoulli, Zip, and Jaz DOS formatted disks can be mounted on a Macintosh.
} The disk needs to be Preformatted or formatted on a PC using a current
} version of either our Win95 or DOS/Win 3.1 utilities (Iomega Tools for
} Win95, Iomega SCSI or Zip Tools). The Macintosh needs to be running our
} driver version 3.5.3 for Bernoulli drives, version 4.2 or 4.3 for the Zip
} drive, or version 4.3 for the Jaz drive.
}
} The Macintosh needs to have a DOS mounting utility. The commonly found
} utilities are:
}
} 1. Apples' PC Exchange version 2.0.2 or greater. It is part of the Apple
} System 7.5 upgrade. This version of PC Exchange can be used with any
} version of System 7 and is available as a standalone product.
} 2. Insignia Solutions Access PC version 3.0 or higher.
} 3. Dayna Communications DOS Mounter Plus version 4.0 or higher.
}
} Each of these software packages will allow the Macintosh to mount DOS
} formatted disks. Consult with your local software vendor for information
} on which package is best suited for your needs.
}
} The Macintosh hierarchal file system (HFS) is not compatible with the FAT
} (DOS style) file system. Because of this we recommend the following
} precautions:
}
} 1. Create the directory structure on a PC and only copy files to the
} disk. Do not create or delete folders on the disk with the Macintosh.
}
} 2. Have backup copies of any file you are transfering. These files should
} be in their orginal format.
}
} 3. Disks that transfer files regularly should be formatted in a PC at
} regular intervals. Once every six months should be sufficient. The more
} often the disk is written to on the Macintosh, the more often this should
} be.
}
} Important Support Note:
} Each of the software packages mentioned should mount Iomega disks, just
} like they would mount a DOS formatted floppy disk. If any usability,
} installation, or compatibility problems are encountered contact the
} software vendor for technical assistance. Iomega does not provide
} technical support for these non-Iomega products.
}
} Problems we have seen:
}
} 1. You insert a DOS formatted disk into the drive and the Mac wants to
} initialize it.
}
} a. You can get this message if another driver is controlling the drive.
} Insert a Mac formatted disk into the drive and then Get Info about the
} disk itself. The line that says Where will tell you what driver is in
} control of the drive. At the end of the line you will see (3.5.3) for
} Bernoulli drives, (4.2) or (4.3) for Zip drives, and (4.3) for Jaz
} drives. Anything else would be conflict.
}
} b. The disk was formatted with Windows NT or a non-standard DOS format.
}
} 2. You try copying a singe large file to the disk and get the message
} "There isn't enough room on the disk."
}
} a. The DOS mounting utility has enough overhead in converting the file
} to limit the size of the file you can copy to the disk. It is roughly
} two-thirds the capacity of the disk.
}
} 3. After copying over about 500 files, the files seem to become corrupt.
}
} a. PC Exchange has a limit of the number of files it can copy just like
} DOS has a limit of the number of files that can be in the root directory.
}
} We have not seen a utility to mount Mac formatted disks on DOS systems.
}
} Iomapple
} Iomega Technical Support

} } } } } } } } } } } }
}
} WHAT ZIP DRIVE MODELS ARE AVAILABLE?
}
} A Zip SCSI model and a Zip Parallel Port model. The SCSI model can be
} used with Macintosh or a PC that is using a ZIP compatible SCSI adapter.
} The SCSI model has two 25 pin SCSI ports (for chaining), a SCSI ID
} switch, and a SCSI termination switch. The Parallel Port model can only
} be used with a PC and connects to the computers Parallel Port.
}
} WHICH SCSI ADAPTERS CAN BE USED WITH THE SCSI ZIP MODEL?
}
} Iomega provides a optional low cost SCSI adapter called the Zip Zoom
} Accelerator. Other ASPI compatible SCSI adapters can be used in
} conjunction with Iomega SCSI driver software.
}
} } } } } } } } } } } }
} Formatting a DOS ZIPs on the Mac (w/ PC exchange installed)?
}
} Yes there is a way of doing this. Go into the Tools program and click on
} Erase Disk. On the next window, select the long, ten minute format. Start
} the format and then stop it after five seconds and quit Tools. This will
} pop the disk out. Reinsert the disk and PC Exchange should take over and
} ask how you would like it formatted.
}
} We have seen problems doing it this way. Sometimes the newly formatted DOS
} disk isn't readable. We suggest that the formatting be done on a DOS
} system.
}
} Iomapple
} Iomega Technical Support

****************************************************************************
******

I hope these clarifications will help using the ZIP diks for image
transfers. For future reference, I have included this info in my web
page. Thanks Klaus



******************************************************************************
* : *
* Klaus-Ruediger Peters, Ph.D. : WWW Home Page: *
* Director, Molecular Imaging Laboratgory : *
* Biomolecular Structure Analysis Center : Molecular Imaging Laboratory *
* University of Connecticut Health Center : http://panda.uchc.edu/ *
* 263 Farmington Ave. : htklaus/index.html *
* Farmington, CT 06030-2017; U.S.A : Differential Hysteresis *
* : Processing Demo at http:// *
* Tel: (203) 679-3977; Fax: (203) 679-1989 : panda.uchc.edu./htbit/indiv/ *
* e-mail: Peters-at-BSAC.UCHC.EDU : /software_docs/dhp.html *
* : *
****************************************************************************
**






From: Owen P. Mills :      opmills-at-mtu.edu
Date: Sun, 10 Mar 1996 13:30:32 -0500
Subject: Need phone number in U.K.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I'm trying to contact the manufacturer of a power supply in our Oxford color
monitor, made in the U.K. Could someone in the U.K. help me by providing an
address & phone number for Cotron Electronics Ltd.?

Thanks.
Owen P. Mills
Michigan Technological University
Metallurgical & Materials Engineering
Rm 512 MME Building
Houghton, MI 49931
906-487-2002
906-487-2934 FAX
opmills-at-mtu.edu





From: Yves Maniette :      yves-at-giga.sct.ub.es
Date: Sun, 10 Mar 1996 15:16:20 +0000 (GMT)
Subject: Re: alignment of CM30+other problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tony,

The complete alignment of the CM30 takes about 1 hour, so if you happen
to have a computer connected to the microscope (allowing you to save and
load alignments data) it might be worth trying the alignment.

By the way I have another question about light variation. On our CM30,
the position of the light varies in a notable way depending on the C2
condensor excitation. If the C2 current is set at maximum, then back to
cross-over we do not get the light at the same place than if we set C2
current at minimum and then back to cross over. (at X 20000 the light
gets nearly out of the screen. I wrote to Philips, they told me that the
microscope was "within specification". I would have liked a more elaborated
answer, or even new inputs from CM microscopes users.

Yves MANIETTE
Universitat de Barcelona
Serveis Cientifico Tecnics
Unitat ESCA TEM
Carrer Luis Sole i Sabaris
E-08028 BARCELONA

Tel +34 3 402 16 95
Fax +34 3 402 13 98

On Fri, 8 Mar 1996, Wharton Sinkler wrote:

}
} Hello all:
}
} I am using a Philips CM30ST for high resolution imaging. I have had a
} problem performing the critical objective lens direct alignment "current
} rotation center". Ideally, one should obtain a stationary point on the
} specimen, rotating about its center at high magnification with a
} low-frequency wobble applied to the objective lens current. I can get
} the stationary point on the specimen reasonable satisfactorily.
} The probjem is that the beam moves fairly significantly when the objective
} lens wobbler is on.
}
} I have performed the coma free direct alignment, which should ensure that
} the beam is travelling along the optic axis of the objective lens. This
} works well. However, it seems to me that if wobbling the objective lens
} current can shift the beam, it can't be travelling the optic axis.
}
} Is there any relatively quick way to correct the problem of beam shift
} with objective lens current change? I can imagine that a complete
} realignment of the microscope might get rid of the problem. However,
} this is both very time-consuming and also not very instructive as to the
} cause of the problem.
}
} Wharton Sinkler
}
} ==================================================
}
} Dr. Wharton Sinkler
} GKSS Forschungszentrum Geesthacht
} Abteilung WA
} D-21502 Geesthacht
} Germany
}
} tel: (49)(4152)872542
} fax: (49)(4152)872534
}




From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Sat, 9 Mar 1996 11:39:23 -0600
Subject: Re: microscope disposal -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02120d02ad677209c345-at-[128.174.23.225]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Another idea: if you know someone with access to a decent machine
shop, and is skilled, you can have the scope column sawn in half
longitudinally to expose the parts _in situ_. We have a couple of TEMs
treated in this way on display, and they are useful teaching tools. (This
was done many years ago by an engineer who used to work here.)

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: Kukin V.N. :      lemi-at-mx.iki.rssi.ru
Date: Sat, 9 Mar 1996 13:00:52 +0300 (MSK)
Subject: Need program

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {199603091507.KAA28839-at-IndyNet.indy.net}
To: Microscopy list {microscopy-at-Sparc5.Microscopy.Com} , SPM List {spm-at-di.com}


I am trying to find any program that will allow to calculate
the isolated islands area and their size distribution using digi-
tal TEM images of metal or metal oxide thin films.

Thanks for any help that may be offered.
Vladimir Kukin,
Electron Microscopy Lab,
Moscow Institute of Electronic Technology (Technical University)
lemi-at-mx.iki.rssi.ru




From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Sat, 9 Mar 1996 09:44:46 +0000 (GMT)
Subject: Re: Lehigh address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In order to get information aboput the excellent SEM and XRMA Short
Courses being offered by Lehigh University contact Ms. ShAron Coe by
e-mail at slc6-at-lehigh.EDU Sharon is a prompt e-mail replyer. See you in
June

Patrick Echlin
University of Cambridge UKOn Wed, 6 Mar 1996, MARK DARUS (216) 266-2895
wrote:

} Could someone send me information, or an E-Mail contact, for
} Lehigh (spelling may be wrong), in PA, so I can get information
} on courses offered.
}
} Thanks
}
} Darus-at-cle.dnet.ge.com
}




From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Fri, 8 Mar 1996 13:38:39 -0500
Subject: Core Facility Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d00ad662e957006-at-[128.206.15.189]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

POSITION AVAILABLE: Associate Director, Molecular Cytology Core Facility
(confocal scanning laser microscopy, low light video microscopy, image
processing/analysis, microtomy, immunocytochemistry and in situ
hybridization). Individual with experience in some or all of these areas
is sought to train users, maintain instruments, develop protocols for a
multiuser facility. PhD desirable but not required for individuals with
extensive experience. Women and minority candidates are especially
encouraged to apply. Address applications (CV and 3 letters of reference)
or inquires to:
Thomas E. Phillips, Ph.D.
Division of Biological Sciences
3 Tucker Hall, University of Missouri
Columbia, MO 65211
573-882-4712
tphillips-at-biosci.mbp.missouri.edu.








From: dshubito-at-d.imap.itd.umich.edu (Dennis Shubitowski)
Date: Fri, 8 Mar 1996 16:43:24 -0500
Subject: Summary: Polaroid Film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02120d01ad66581f3765-at-[141.211.157.94]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Many thanks to all who responded about my Polaroid film dilemna. The emulsion
ended up washing right off in an overnight soak. Hopefully this does not
happen again. If it does, however, I am armed with an arsenal of tips!


Thanks again,

Dennis Shubitowski
University of Michigan
School of Dentistry


Here is a summary of the responses:

I have never been successful in separating T55 negatives that have dried
together.

John
chandler-at-lamar.ColoState.EDU

-----

I did SEM for 7 years at Wayne State and ran into that problem at least several
times. Let me tell you straight: it can't be done and isn't worth the effort.
You'll always get some junk to remain on the negative ruining it forever. Oh, I
hope the negatives aren't too important!

Walt Bobrowski
Parke-Davis Research

-----

Have you tried a little detergent in the water? The liquid type is the
best to use.

I don't know if this is a sure thing, I always used a water soak, but my
success may be due to the local water.

Hope this helps.

Ellie Solit
The Microscope Book

-----

I have used a AGFA AGEPON-WATER SOAK on TEM nagatives before
That worked

Stephan H Coetee
Electron Microscope Unit
Private Bag 3
Wits
2050 Stephan-at-Gecko.biol.WITS.ac.za

-----

Yes, I have successfully unstuck pairs of Polaroid negs. Actually, the
water soak will work, but you have to soak them for a long time, at least
overnight, say 12-16 hours. Then you have to slowwwwwwwly pull them apart to
avoid pulling the emulsion off. A "pucker" mark may remain where they were
joined, but usually its translucent so that a positive print will seldom reveal
the pucker mark, especially on "busy" image detail, and as long as original
image density was not disturbed.

When this happens to my negs, its usually in a spot no larger than the size of a
quarter (25 cent piece). If its most of the area of the negs, that is serious
trouble.

When I dry Polaroid negs in racks, I skip slots so that there is lots of space
between adjacent negs to prevent sticking from occuring as the negs "wiggle" as
they air dry.

Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu

-----

I have sometimes (!) been successful by giving the negatives a prolonged
soak (days) in water. the problem, of course, is that the emulsion can be
washed off - you have to hope that the negatives unstick before you lose the
emulsion. Good luck!

Tony Garratt-Reed

-----

I remember a water soak working for me, but not without waiting over a
weekend. Even then it left a patch ...

{\/} /\ {\/} /\ {\/} /\ {\/} cognito, ergo zZOooOM {\/} /\ {\/} /\ {\/} /\ {\/}
Michael Shaffer, R.A. - University of Oregon Electron Probe Facility
mshaf-at-oregon.uoregon.edu -or- mshaf-at-darkwing.uoregon.edu
http://darkwing.uoregon.edu/~mshaf/epmahome/

-----

I have had limited success with an overnight Photoflo soak
followed by ultrasonication. Films stuck emulsion to film
back seem to separate easier than those stuck emulsion to
emulsion.

richard.mount-at-mailhub.sickkids.on.ca

-----

I agree with Richard, above, that film back stuck to emulsion is easier to
seperate, and this is the case I usually run into as I dry my Polaroid negs in
racks with the emulsion of one facing the back of the next neg.

A point that I left out in my previous note describing successful use of
overnight water soak to seperate stuck negs is how I set this up. As with many
techniques, success lies in the details of the method. I load a pair of stuck
negs into my usual neg drying rack with edges of one neg in one set of slots,
edges of other neg in adjacent set of slots. The effect is to separate the negs
where they are not stuck to enable water to get directly to the stuck area.
Success depends on getting good water soak and softening of the film emulsion.
If you were to simple lay the two stuck negs flat in a pan of water, might not
get as good a soak as when the films are seperated a bit to allow water access.

I hope this helps.

Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu

-----

YES, virginia, polaroid negatives can usually be unstuck, just soak in
very hot water about 10 -20 minutes, then GENTLY pull them apart, GRADUALLY
turning and pulling a little untilthe whole gelatine that is stuck gets
soaked through and losened. Some times if more completely dried out, the
emulsion of one will be stuck to the second, usually the second can be saved
completely, but some of the first will be lost (natually the center, best
part of the image)
good luck

Alan S. Pooley ,PhD Bivalve shell SEM & shape analysis
SEM/Morphometrics lab
Marine & Coastal Sciences
Rutgers University
908 932 8959 ext 225
Pooley-at-ahab.rutgers.edu







From: mager-at-unixg.ubc.ca (Mary Mager)
Date: Fri, 8 Mar 1996 20:14:46 -0800
Subject: Re: Microsc.Soc Canada Web page

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Dear Mary, I read your notice about Stan klosevych's papers in the MSC
} bulletin,
} and tried to locate the NSC home page on the web address you gave. My computer
} running under Netscape did not recognise this. i would still like to know more
} of Stan's articles, as i teach microscopy. Can you help. TIA Jeremy Sanderson.
Dear Jeremy,
As several of my friends informed me, the Web site I gave for the MSC had a
small typo that made it unusable. In my defense, I copied it correctly from
the current MSC Bulletin. They had it wrong. The correct address is:
http://gause.biology.ualberta.ca/craig.hp/MSC.SMC/Mic.Soc.hp
As I said, there is also a link from the MSA WWW site.
Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Eng, UBC
309 - 6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648, fax: 604-822-3619
email: mager-at-unixg.ubc.ca






From: Rudi Rottenfusser :      rrotte-at-mbl.edu
Date: Fri, 8 Mar 1996 12:00:42 -0500
Subject: Zeiss Parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603081701.MAA04205-at-hoh.mbl.edu}
Sender: {rrotte-at-mbl.edu}
microscopy-at-Sparc5.Microscopy.Com
Priority: normal
X-mailer: Pegasus Mail/Windows (v1.22)

Good morning.

For your Universal M microscope equipped with the "Vertical
Illuminator IIC", you will need a 22mm diameter glass plate (semi-
transparent mirror). You can install this into your "H-PL" housing.

It is still available as a "spare part" from Zeiss under catalog
number 466260-0010 at U.S. $72.00. For ordering spare parts,
please call
U.S.: 800/233-2343
Canada: 800/387-8037

The other option you always have is to call one of the companies
which manufacture all kinds of filters such as

Chroma Technologies (Vermont): Tel. 802/257-1800
Omega Optical (Vermont) Tel. 802/254-2690.

Please e-mail me if I can be of further help.



You wrote " I have a Zeiss Universal 'M' microscope which is missing a few
} parts, which are now VERY (?) expensive (from Zeiss).
}
} I require a REFLECTOR for a 'Vertical Illuminator IIC'; either a
} H-PL, or H-PR will do. Only the reflector is required as I have the
} housing. Does anyone have one of these ? Or does anyone know if it is
} possible to 'make' one using a 'double sided semi-transparent mirror' ?
} ==============================================================================
} Steve Schwarz sschwarz-at-morgan.ucs.mun.ca
} Dept. of Earth Sciences
} Memorial University of Newfoundland
} Newfoundland
} CANADA
} A1B 3X5
} 1-709-737-8142
} -737-2589 FAX





From: Walter Arno Mannheimer :      wamann-at-metalmat.ufrj.br
Date: Fri, 8 Mar 1996 14:09:22 EST3EDT
Subject: disposal of old microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We decomissioned our Stereoscan Mk II after 25 years of service, as
an instrument with open (graduate) student access. Managing the
electronic tubes just became impossible.
The Engineering School mantains a Museum, were we gave it honourable
burial. But first we took out pumps, oscilloscope, camera, some
gauges, etc. which continue to serve usefully. The empty shell still looks nice,
and who would know the difference.
It is now among scale models of locomotives, our first computer, etc.
This instrument was installed in 1969. We know it was the first SEM in
Brazil, probably also in South America, and we like to think it might
have been the first in the Southern Hemisphere. Anyone like to
comment on this?
Prof.Walter A.Mannheimer
Metallurgy and Materials Engineering
Federal University of Rio de Janeiro
POBox 68505
21945 Rio de Janeiro Brazil
Voice 5521 280-7443 (Dept.office) 5521 590-0579 (direct)
Fax 5521 290-6626 Email: wamann-at-metalmat.ufrj.br




From: Theresa A. Fassel :      fassel-at-post.its.mcw.edu
Date: Fri, 8 Mar 1996 17:00:05 -0600 (CST)
Subject: fungi samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi

I will receive Candida albicans samples in my lab for service. I am
looking for a good preservation of cell wall, cell membranes, nuclei, mitos,
ribosomes, etc. I must keep the cell wall intact for this study. ALso,
I want to avoid acrolein.

Does anyone have a solution ?? ANy advise?


Thanks in advance

Theresa

Dr. Theresa A. Fassel
Sr. Research Associate fassel-at-post.its.mcw.edu
Department of Microbiology (414)-456-8410
Medical College of Wisconsin Fax (414)-266-8522
8701 Watertown Plank Road
Milwaukee, WI 53226-0509






From: lmiller-at-ux1.cso.uiuc.edu (Lou Ann Miller)
Date: Sat, 9 Mar 1996 08:51:29 -0500
Subject: Re: fungi samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510100ad67357c7777-at-[128.174.23.58]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi Teresa,



I have done Yeast, lichen, soybean and tobacco seedlings with microwave
fixation and processing, and have been quite pleased with the vast
improvement in results.

I have used both a standard microwave and a nice one from Ted Pella
specifically for EM.

There are many ways to do this, but the following is what we follow in our lab:

Microwave rules:
1. Always use at least a 250 ml water load to collect the heat
and keep it off your sample.
I change this every 16 seconds of use ( 2 fixations). ****
Always set the water load in
the same place each and every time.

2. Calibrate the Microwave. Set the water load/ beaker in
it's designated spot.
Use liquid crystal sheets and microwave about 15 seconds.
-- Observe where the hot spots show up, and avoid them.
Pick a place preferable
inbetween hot spots.

3. Always do a pilot study first.

4. Things to avoid like the plague:

--- Always chill your sample in wetted down crushed ice for
at least 30-60 seconds
before microwaving. This will allow MW, but keep heat off
your sample.
**** As soon as you are finished microwaving , put back
into the ice for a moment.

---- Always wipe off the outside of your vial, for water
draws heat ( hence the
water load situations) Also never allow water to be
anywhere but in the
waterload beaker.

--- Using the same principle, only use enough fixative/
solution to just cover the
sample. I usually use about 1-2 mm high. Greater volumes
will EXPONENTIALY
(sp) increase the heat load to your sample. The difference
between 6 drops and 1 ml
can be 40C!!!!!

Processing times:
==============


You have to experiment with this.

I'd start out with possibly the following for things with thick walls and
for plants:

1. Ice sample, wipe, microwave for 7 seconds, ice.

2. Repeat the above for a total of 3 times.

3. Rinse in buffer.

4. Put into OsO4---2%aq.

Ice, MW 7 seconds, let set 10 min................change solution
and repeat once more.
And save the last OsO4 on the sample.

****** it is amazing how MW speeds up processing times, for the
first time ever after
starting microwave, I have to really watch OVER staining enbloc.
So...... repeat a 3rd time if you don't like the staining of the
first, but try doing only 2
OsO4's at first.

5. To the last OsO4 solution on sample, Add equal volume of 3%
Potassium Ferricynide.
Set for 20 minutes only.

6. Rinse with water.

7. Add filtered saturated aq Uranyl Acetate, no older than 10 days.
( Might be our water problem)

Ice, MW 7 seconds Ice, and let set for 15 minutes at room temperature.

REPEAT

8. Dehydrate as norm. 8-10 min. each step.

9. Infiltration: -- these will get warm, it's OK, and even helps greatly.


1:1 MW 10 seconds, no ice, incubate (rotating)15 minutes and REPEAT

1:4 MW 10 seconds, no ice, incubate 1 hour and repeat.

Pure epoxy: MW 10 seconds, no ice Incubate 1 hour intervals and
overnight.
Repeat MW in am, make at least 2 changes of
epoxy.

** Also helps to spend overnight in vacuum in a
desicator hooked up to
house vacuum.

The difference I've seen in structure, staining and non-pitting out of the
specimens vs the more traditional methods I've used is incredible. We like
microwave so much we now use it for all of our samples.

I'm going to be leaving for a week, but if you'd like to know more, I can
get you references etc when I'm at work when I get back. Our methods are
based on Logan's, Ted Pella's and my 2 years of working through techniques.


Good Luck,

Lou Ann



} Hi
}
} I will receive Candida albicans samples in my lab for service. I am
} looking for a good preservation of cell wall, cell membranes, nuclei, mitos,
} ribosomes, etc. I must keep the cell wall intact for this study. ALso,
} I want to avoid acrolein.
}
} Does anyone have a solution ?? ANy advise?
}
}
} Thanks in advance
}
} Theresa
}
} Dr. Theresa A. Fassel
} Sr. Research Associate fassel-at-post.its.mcw.edu
} Department of Microbiology (414)-456-8410
} Medical College of Wisconsin Fax (414)-266-8522
} 8701 Watertown Plank Road
} Milwaukee, WI 53226-0509

***************************
Lou Ann Miller
Microscopic Imaging Lab
College of Vet. Medicine
University of Illinois
2001 S Lincoln Ave
Urbana,Illinois 61801
217-244-1566
lmiller-at-ux1.cso.uiuc.edu

Microscopy Home Page:
http://www.cvm.uiuc.edu/MicImagLab/MicImagLab.html

Personal Home Page:
http://www.cvm.uiuc.edu/HomePages/LouAnnMiller/LAM.html






From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 11 Mar 1996 17:11:33 -0400
Subject: RE-Origin of C Particles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1385558009.21007-at-mse.engin.umich.edu}

Subject: Time: 5:05 PM
OFFICE MEMO RE:Origin of C Particles Date: 3/11/96

It may well be that by carrying out a proper analysis of your materials
using the various technidques of optical microscopy (polarized light,
refractive index meas, etc.), you can get info about the source of the carbon
particles. The people at McCrone Associates Inc (850 Pasquinelli Dr,
Westmont, IL 60559 (Ph: 708-887-7100, Fx: 708-887-7417) are world reknown
specialists in identifying particulate materials and determining their
sources. I suggest you get in contact with them.
W. C. Bigelow (bigelow-at-umich.edu)





From: Johari, Dr. Om :      73211.647-at-CompuServe.COM
Date: 11 Mar 96 08:09:34 EST
Subject: Updated info on SPM program at Bethesda

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"Margoninski, Yohanan" {Ymargo-at-VMS.HUJI.ac.il} ,
"Markewicz, Peter" {pmarkiew-at-chem.utoronto.ca} ,
"van der Heide, Paul" {ldm-at-apollo.numis.nwu.edu} ,
"Marti, Othmer J." {Othmar.marti-at-physik.uni-ulm.de} ,
"Maugel, Tim" {maugel-at-zool.umd.edu} ,
Microscopy {Microscopy-at-aaem.amc.anl.gov} ,
"Mochizuki, Yasunori" {mochizuk-at-sci.cl.nec.co.jp} ,
"Molecular Imaging (Gatan)" {info-at-molec.com}

-- [ From: Dr. Om Johari * EMC.Ver #2.0 ] --
Expanded recipient data:
To: Ichikawa, Masakazu \ Internet: (ichikawa-at-jrcat.or.jp)
To: INTERNET:SInto-at-aol.com \ America On-Line: (sinto)
To: INTERNET:jak-at-zurich.ibm.com \ Internet: (jak-at-zurich.ibm.com)
To: Johnson, Robert E. \ Internet: (rej-at-virginia.edu)
To: DAVID JOY \ Internet: (djoy-at-utk.edu)
To: Keller, David \ Internet: (dkeller-at-triton.unm.edu)
To: Kavanagh, Karen \ Internet: (kkavanagh-at-ucsd.edu)
To: Kenny, Peter \ Internet: (p.g.kenny-at-bradford.ac.uk)
To: Kenny, Peter \ Internet: (p.g.kenny-at-bradford.ac.uk)
To: Kiyoshi Kobayashi \ Internet: (kiyoshi_kobayashi-at-vnet.ibm.com)
To: Bernd W. Koenig \ Internet: (bernd-at-helix.nih.gov)
To: Kotorman, Louis \ Internet: (kotorman-at-craft.camp.clarkson.edu)
To: Ondrej Krivanek \ CompuServe: (73752,3103)
To: Kruit, Pieter \ Internet: (kruit-at-cpo.tn.tudelft.nl)
To: Kundmann, Michael K. \ CompuServe: (73752,3070)
To: Lagally, Max G. \ Internet: (lagally-at-engr.wisc.edu)
To: INTERNET:lang-at-watson.ibm.com \ Internet: (lang-at-watson.ibm.com)
To: Lazebnik, Yuri \ Internet: (lazebnik-at-cshl.org)
To: Lea, Peter J. \ Internet: (spectral-at-inforamp.net)
To: Leapman, Richard D. \ Internet: (leapman-at-helix.nih.gov)
To: LEO (UK) \ Internet: (info-at-leo-em.co.uk)
To: Lewis, Nathan S. \ Internet: (nslewis-at-juliet.caltech.edu)
To: HaLi \ Internet: (hl-at-padns1.phy.tu-dresden.de)
To: Lieber, Charles M. \ Internet: (cml-at-cmliris.harvard.edu)
To: Lindsay, Stuart \ Internet: (stuart.lindsay-at-asu.edu)
To: Lyo, In-Whan \ Internet: (lyo-at-watson.ibm.com)
To: Marek Malecki \ Internet: (malecki-at-macc.wisc.edu)
To: Murphy, Judy \ Internet: (murphy-at-ms.sjdccd.cc.ca.us)
To: Nagy, Peter \ Internet: (p.nagy-at-r1.atki.kfki.hu)
To: Newbury, Dale E. \ Internet: (newbury-at-enh.nist.gov)
To: INTERNET:dmnewns-at-watson.ibm.com \ Internet: (dmnewns-at-watson.ibm.com)
To: Nicholov, Rina \ Internet: (nicholov-at-botany.utoronto.ca)
To: Jun Nogami \ Internet: (nogami-at-csd.uwm.edu)
To: Norde,Willem \ Internet: (willem-at-metten.fenk.wau.nl)
To: Peter J.A. Nordlander \ Internet: (nordland-at-kaitum.rice.edu)

please circulate/post

Please complete/return form below to remain/get on our lists.

Scanning Microscopy International
P.O. Box 66507, Chicago (A.M.F. O'Hare), IL 60666-0507, U.S.A.
Telephone: (708) 529-6677 / FAX: (708) 980-6698
E.mail: 73211.647-at-compuserve.com

Scanning Microscopy 1996 meeting
May 11-16, 1996, Bethesda, Maryland (suburb of Washington, DC)

Symposium on: Scanning Probe Microscopies and Related Techniques
for the Biological and Materials Sciences

Note: Nearly 35 SPM related papers will be presented during the
program on "Pattern Formation and Nanoscaled Structures in Thin
Film Formation" at the same time at same venue. THOSE PAPERS ARE
SEPARATELY LISTED BELOW.

Program organizers: Dr. David P. Allison, Oak Ridge Natl. Lab., TN
(E-mail: allisondp-at-ornl.gov); Prof. Chunli Bai, Chinese Acad. Sci.,
Beijing (E.mail: clbai-at-infoc3.icas.ac.cn); Prof. Masamichi
Fujihira, Univ. Yokohama, Japan (E.mail:
mfujihira-at-bio.titech.ac.jp); Dr.
Heinrich J.K. Hoerber, European Molec. Biol. Lab., Heidelberg,
Germany (E.mail: hoerber-at-embl-heidelberg.de); and Prof. Douglas J.
Thomson, Univ. Manitoba, Winnipeg, Canada (E.mail:
thomson-at-ee.umanitoba.ca).

VERY FEW papers can still be offered: please contact one of the
organizers or Dr. Om Johari at Scanning Microscopy International.

The registration fee for the entire 6-day conference (without
subscription to any of the SMI journals' 1996 volumes) is either
$100 (if payment reaches SMI before March 31, 1996) or $120 (from
April 1, 1996); the fees with 1996 subscription to one, two or all
three SMI journals (Scanning Microscopy, Cells and Materials, and
Food Structure) are respectively: $180, $220, or $270 (for US
delivery addresses), or $210, $270, or $340 (for outside US
addresses). The room rates (exclusive of applicable taxes,
currently 15%) at the Hyatt Regency Bethesda (One Bethesda Metro,
Bethesda, MD 20814, USA; phone: 301 657 1234; FAX 301 657 6453)
are: Single (1 person - 1 bed): $100; Double (2 persons - 1 bed)
or Twin (2 persons - 2 beds): $110. Please make room reservations
directly with the hotel.

List of presentations (as of March 10, 1996) in alphabetical
order (the name of the presenter is capitalized when there are more
than one authors)

M.J. ALLEN, Digital Instruments, Santa Barbara, CA; E.M. Bradbury,
R. Balhorn:
The Chromatin Structure of Well-Spread Demembranated Human Sperm
Nuclei Revealed by AFM

D.P. ALLISON, P.S. Kerper, M.J. Doktycz, T. Thundat, R.J. Warmack,
Oak Ridge National Lab., TN:
High Resolution Physical Mapping of EcoRI Restriction Sites on
Intact Cosmids by AFM Imaging

P.C. Zhang, C. BAI, P.K.H. Ho, Q. Li, Y. Dai, Y.S. Wu, B.F. Sheng,
Chinese Acad. Sci., Beijing, China:
AFM Study of Interactions Between Tumor Necrosis Factor and Its
Monoclonal Antibodies

A.A. BERGMAN, A.P. Quist, C.T. Reimann, S.O. Oscarsson, B.U.R.
Sundqvist, Uppsala Univ., Sweden:
Antibody-Antigen Docking Observed with Tapping-Mode Scanning Force
Microscopy in Air

A.A. BUKHARAEV, N.V. Berdunov, D.V. Ovchinnikov, Kazan Physical-
Technical Inst., Russia:
Three-Dimensional Probe and Surface Reconstruction for Atomic Force
Microscopy Using Deconvolution Algorithm

R.J. Pylkki, G.J. COLLINS, Topomatrix, Santa Clara, CA:

Scanning Thermal Microscopy with a Positive Probe

E.D. DAHLBERG, R. Proksch, S. Foss, G. Skidmore, C. Merton, Univ.
Minn, Minneapolis:
Review: Magnetic Force Microscopy and Magnetotactic Bacteria

P. DEWOLF, T. Trenkler, T. Clarysse, W. Vandervorst, IMEC, J.
Snauwaert, L. Hellemans, Leuven, Belgium:
Electrical Characterization of Submicrometer Silicon Devices by
Cross-Sectional Contact Mode AFM (to be presented in the
Semiconductor program)

S.J. EPPELL, R.E. Marchant, Case Western Reserve Univ., Cleveland,
OH:
Atomic Force Microscopy of Bone: More than an Image (to be
presented during the Bone Biology program)

R. ESCHRICH, Max-Planck-Inst. Biochem., Martinsried, Germany; G.L.
Kumar, T.A. Keil, R. Guckenberger:
AFM on the Olfactory Dendrites of Giant Silkmoth Antheraea

G.L. Kumar, Max-Planck-Inst. Verhaltenphysiol., Seewiesen, Germany;
R. ESCHRICH, R. Guckenberger, T.A. Keil:
In Search for Putative Pheromone Receptors on the Membrane of
Olfactory Dendrites in Silkmoths (A. polyphemus and A. pernyi)
Using the AFM and SEM

M. Fujihira, Tokyo Inst. Tech., Japan:
Review: AFM of Solid Surfaces in Aqueous Solutions

L.A. GHEBER, J. Hwang, M. Edidin, Johns Hopkins Univ., Baltimore,
MD:
Imaging of Rough Biological Samples in Liquid with Fluorescence
Near Field Scanning Optical Microscopy

D.T. GODDARD, A. Steele, I.B. Beech, British Nuclear Fuels,
Preston, UK:
AFM of Bacterial Biofilms with Application to Microbially
Influenced Corrosion

H. Hansma, Univ. Calif., Santa Barbara:
Review: Atomic Force Microscopy of Biomaterials

M. HARA, W. Knoll, RIKEN, Saitama, Japan:
Review: STM and AFM Studies of Self-Assembled Monolayer Growth

D.O. HENDERSON, Y.S. Tung, R. Mu, Fisk Univ., Nashville, TN; W.A.
Curby, M. Mercado, Fed. Aviation Rech. Ctr., Atlantic City, NJ:
Optical and Atomic Force Microscopy of Pentaerythritol Tetranitrate
Nanoclusters on Si(100)

C.L. Mosher, E. HENDERSON, Iowa State Univ., Ames:
Review: Chemical and Biomolecular Force Detection

R. Eckert, S. Jeney, W. Oeffner, J.K.H. HOERBER, European Molecular
Biology Lab., Heidelberg, Germany:
Measuring Surface Forces with the AFM

D.H. HUANG, Y. Yamamoto, Yamamoto Quantum Fluctuation Project,
Tokyo, Japan:
Hydrogen Atom Extraction and Redeposition on Hydrogen-Terminated
Silicon Surface with Ultra High Vacuum STM at Room Temperature

A. IKAI, Tokyo Inst. Tech., Japan, K. Mitsui, M. Hara:
Review: Measurements of Mechanical Parameters of Proteins and
Chromosomes

A. IKAI, H. Takeuchi, S. Kawauchi, Tokyo Inst. Technol., Japan
STM Study of Stearoyl Amide and Anilide

M.D. JOHNSON, Univ. Oklahoma, Norman, and H.W.M. Salemink:
Review: Cross-Sectional STM on Semiconductor Heterostructures (to
be presented during the Semiconductors program)

I. KARL, J. Bereiter-Hahn, Univ. Frankfurt, Germany:
Regulation of All Surface Motility Revealed by Scanning
Acoustic Microscopy

L. KUUTTI, J. Pere, J. Peltonen, O. Teleman, VTT Biotech. Food
Res., Espoo, Finland:
Interpretation of High Resolution AFM Images of Crystalline
Cellulose Using Molecular Modelling

R.M. LeBlanc, Univ. Miami, Coral Gables, FL:
Review: Scanning Probe Microscopies of Langmuir-Blodgett Films

G. LEE, A.D. MacKerell, L.A. Chrisey, R.J. Colton, US Naval Res.
Lab., Washington, DC:
Measuring Inter- and Intramolecular Forces in Biomolecules

A.L. LITVIN et al., US Army Natick R&D Center, Natick, MA
AFM and Optical Studies of Organic Thin Films with Hydrogen-Bonded
Networks

L.S. Shlyakhtenko, S. Adhya, T. Aki, Y. LYUBCHENKO, Arizona State
Univ., Tempe:
AFM Studies of Regulatory Nucleoprotein Complexes

F.K. MANTE, Univ. Pennsylvania Dental School, Philadelphia, G.
Baran:
Nanoindentation Studies of Single Crystal Titanium

J.F. Marchiando, NIST, Gaithersburg, MD::
Methods for Interpreting Measurements from a Scanning Capacitance
Microscope (to be presented during the Semiconductors program)

M.T. MCDERMOTT, Univ. Alberta, Edmonton, Canada, J.-B.D. Green,
M.D. Porter:
Review: Chemical Mapping with Force Microscopy

M. Rivera, M. MILES, R.L. Williamson, Univ. Bristol, UK:
Phase Transitions in Liquid Crystals Observed by SPM

W. MIZUTANI, M. Motomatsu, H. Ogiso, H. Tokumoto, Nat. Inst. Adv.
Interdiscpl. Res., Tsukuba, Japan:
STM with Local Non-Linearity Detection of Organic Thin Films and
Ion Irradiated Graphite

D.J. MUELLER, F. Schabert, A. Engel, Univ. Basel, Switzerland:
Review: Structural Changes of Native Membrane Proteins Monitored at
Subnanometer Resolution with the AFM

F. MUELLER, A.-D. Mueller, Tech. Univ., Chemnitz, Germany:
Scanning Capitance Microscopy of Composite Materials

H. MURAMATSU, Seiko Instruments, Chiba, Japan; T. Ataka, N. Chiba,
K. Nakajima, M. Fujihira
Review: Fluorescence Imaging and Spectroscopy of Biomaterials in
Air and Liquid by SNOM/AFM

R. Perez, Univ. Cambridge, UK:
First Principles Simulations of Atomic Resolution in Non-Contact
AFM (to be presented during the Fundamental Physics Program)

R. RAITERI, Univ. Genoa, Italy, H.-J. Butt, Max-Planck-Inst.
Biophy., Frankfurt, Germany:
Changes in Surface Stress Measured with an AFM

B. SAMORI, A. Gordano, I. Muzzalupo, Univ. Bologna, Italy:
DNA Deposition on Mica for Scanning Force Microscopy Imaging

W. Haiss, J.K. SASS, Fritz-Haber-Inst., Berlin, Germany:
STM Surface Stress Measurements in Electrochemistry (tentative)

A.J. Simon, Merck Res. Lab., West Point, PA:
Review: Molecularly Combed DNA: First Stretched, Then Imaged

R.P. Singh, Indian Min. Sci. Tech., New Delhi:
STM and Application Possibilities

Y. Shirane, Univ. Tokushima, Japan:
Surface Observation of Calcium Oxalate Monohydrate Crystals by FE-
SEM and AFM (to be presented during the Stones and Crystals
program)

F. El Feninat, C. Poulin, T. Ellis, E. Sacher, I. STANGEL, McGill
Univ., Montreal, Canada:
The Applications and Limitations of AFM to Complex Biomaterials
Surfaces: Effects of Acid Demineralization on Human Dentin

M.S. UNLU, B.B. Goldberg, Boston Univ., MA:
Review: Characterization of Materials and Devices by Near Field
Optical Scanning Microscopy (to be presented during the
Semiconductors program)

J. Vesenka, Calif. State Univ., Fresno:
Review: The Diameter of Duplex and Quadruplex DNA Measured by SPM

VU THIEN BINH, F. Feschet, V. Semet, S.T. Purcell, Univ. Lyon,
France:
Fresnel Projection Microscopy: Theory and Experiment: Electron
Microscopy with Nanometer Resolution at ~ 200 eV

C. Lebreton, Z.Z. WANG, CNRS Lab. Microstr. Microelec., Bagneux,
France:
Pulse Width Dependence of Nanowriting

G. Brown, M.B. WEIMER, Texas A&M Univ., College Station, TX:
Adsorption and Interaction of Ammonia Molecules on GaAs(110)
Studied with STM

F. Zenhausern, IBM Watson Research Ctr., Yorktown Hts., NY:
New Developments in Near Field Optical Microscopy/Spectroscopy

---

Please circulate / Publicize

Scanning Microscopy International
Post Office Box 66507, Chicago (A.M.F. O'Hare), IL 60666-0507,
U.S.A.
Telephone: (708) 529-6677 / FAX: (708) 980-6698
e.mail: 73211.647-at-compuserve.com

Scanning Microscopy 1996 meeting; May 11-16, 1996, Bethesda,
Maryland

Program on:

Pattern Formation and Nanoscaled Structures in Thin Film Formation

Organized by: Prof. Martin Zinke-Allmang, Univ. Western Ontario,
London, Canada (E-mail: M.Zinke-at-uwo.ca; FAX 519 661 2033), Prof.
Hiroshi Iwasaki, Osaka, Japan (E-mail:
iwasaki-at-sanken.osaka-u.ac.jp; FAX: 81-6-8798404) and Prof. Ellen D.
Williams, Univ. Maryland, College Park (E-mail:
williams-at-surface.umd.edu; FAX: 301 314 9465).

Program Focus: (1) Surface and Interface Defect Formation and
Structure; (2) Phase Separation on Surfaces and Epitaxial Growth
Modes; (3) Surface Roughness, Step Formation, and Dynamics During
Growth; (4) Characterization Techniques with a Focus on Electron
Microscopy and Tunneling Methods; (5) Synthetic Sublithographic
Nanostructures; and (6) Defect Engineering and Related Applications
for Mismatched Materials.

A FEW Papers can still be offered, please contact one of the
organizers or Dr. Om Johari at Scanning Microscopy International.

List of presentations (as of March 11, 1996) in alphabetical order

P. Campbell, Naval Res. Lab., Washington DC:
Nanofabrication with Proximal Probes

G. Carlow, Univ. West. Ontario, London, Canada:
Hummock Formation and Evolution During Ion Bombardment of
Rotating Substrates

O. Enea, Univ. Poiters, France:
Topographic Studies of TiO2 and SnO2 Ceramics by Environmental
Scanning Electron Microscope, SEM and AFM

R. Feenstra, Carnegie Mellon Univ., Pittsburgh, PA:
STM Studies of Semiconductor Thin Film Structures

E.A. Fitzgerald, MIT, Cambridge, MA:
CVD Grown Si-based Semiconductor Film Structures

D. Gerthsen, Univ. Karlsruhe, Germany:
Atomic Scale Investigation of the Mismatch Relaxation in
In0.6Ga0.4As/GaAs(100)

M. Grant, McGill Univ., Montreal, Canada:
Theoretical Studies of the Dynamics of Nanostructure Formation

M. Hong, AT&T Bell Labs., Murray Hill, NJ:
Novel Heterostructures Fabricated Using In-Situ Molecular Beam
Epitaxy

J.W.P. HSU, Q. Xu, Univ. Virginia, Charlottesville, E.A.
Fitzgerald, Y.H. Xie, P.J. Silverman:
Scanning Probe Microscopy Studies of Electrically Active Defects
in Lattice Mismatched Films

R. Hull, Univ. Virginia, Charlottesville:
Real-Time Nanostructural Studies of Defects and Degradation Modes
in Semiconductor Materials and Devices

M. Ichikawa, JRCAT, Tsukuba, Japan:
Observation and Formation of Nanostructures Using a Combined
Microscope with Scanning Reflection and Scanning Interference
Electron Microscopes and STM

A. Iwamoto, T. Yoshinobu, H. IWASAKI, Osaka Univ., Japan:
Scanning Probe Microscopic Studies of Surface Kinetic Roughening

H.-C. JEONG, J.D. Weeks, Univ. Maryland, College Park:
The Dynamics of Steps on Vicinal Surfaces During
Reconstruction-Driven Faceting

H. KIRK, Z. Radzimski, G. Rozgonyi, North Carolina State Univ.,
Raleigh:
Electron Beam-Induced Current (EBIC) Characterization of Defects
in SiO2 and SiO2/Silicon Interface

A. Kowal, Polish Acad. Sci., Krakow, Poland:
Properties of Nickel Hydroxide Layer Formed Electrochemically on
the Surface of Polycrystalline Ni in 1M KOH Studied by In Situ
Atomic Force Microscopy

M. KUNITAKE, N. Batina, K. Itaya, JRDC/ERATO: Itaya
Electrochemistry Project, Sendai, Japan:
In Situ STM Study of Self-Organized Molecular Layers on an
Iodine-Modified Au(111) Surface

J.W. Lyding, Univ. Illinois, Urbana:
UHV-STM Nanofabrication and Semiconductor Interface
Characterization

S. MATLIS, H. Ron, I. Rubinstein, Weizmann Inst. Sci., Rehovot,
Israel:
AFM Study of Gold Oxide Stabilized with Self-Assembled Monolayer

K. Matsumoto, Electrotechnical Lab., Tsukuba, Japan:
Application of STM/AFM Nano-Oxidation Processes to Room
Temperature Operated Single Electron Transistors and Other Devices

J.J. McClelland, NIST, Gaithersburg, MD:
Nanofabrication via Laser Focusing of Atoms

I. Nevernov, Technobiochip, Marcina, Italy:
Sub-Micron Machining of Al Tracks Deposited onto a Silicon Wafer

R. Nishitani, T. Umeno, A. Kasuya, Y. Nishina, Kyushu Inst.
Technol., Iizuka, Japan:
Correlation Between STM-Induced Photon Mapping and the STM
Topography of Nanometer-Size Metal Particles

D.D. PEROVIC, B. Bahierathan, Univ. Toronto, Canada, D.C. Houghton,
H. Lafontaine, J.-M. Baribeau:
Morphological Instability Phase Diagrams: Mapping Thin Film
Strain Relaxation Behaviour

C. K. Shih, Univ. Texas, Austin:
Scanning Probe Microscopy and Spectroscopy of Semiconductor
Heterostructures

J. Stroscio, NIST, Gaithersburg, MD:
Iron and Chromium Growth Studies

I. TERESHKO, V. Khodyrev, E. Lipsky, Mechanical Engg. Inst.,
Mogilev, Belarus:
Self-Organizing Processes and Modification of Thin Films by
Scanning Microscopy

R. Tromp, IBM, Yorktown Hts., NY:
LEEM Related Studies and Optical Microscopy of Semiconductors

L. Whitman, Naval Res. Lab., Washington, DC:
The Atomic Scale Structure of Semiconductor Surfaces: What Makes
the Atoms Smart?

S.-L. YAU, Y.-G. Kim, K. Itaya, JRDC/ERATO: Itaya Electrochemistry
Project, Sendai, Japan:
Scanning Tunneling Microscopy of Chemisorbed Benzene on Rh(111)
and Pt(111) Electrodes in Hydrofluoric Acid Solution

J. ZEGENHAGEN, Max-Planck-Inst. Festkoerperforschung, Stuttgart,
Germany, P. Lyman, M.J. Bedzyk, M. Boehringer:
Review: Analysis of Microscopic Structure of Stress Induced
Reconstructions on Semiconductor Surfaces

M. ZINKE-ALLMANG, T.D. Lowes, Univ. Western Ontario, London,
Canada:
Cobalt Clustering on Si at Low Temperatures

---

Other related programs at the same venue (*: Fliers available):

*Fundamental Physics in Microscopy and Microanalysis (~ 25 papers),

*Scanning Microscopy and Semiconductors: Metrology and Diagnostics
(25+ papers);

Several Biological programs including: Microanalysis and Imaging
(12 papers),
Immunolabelling (8 papers), *Radiation Effects / Apoptosis (25+
papers), *Dentistry (20+ papers), Corrosion Casting (including a
program on Lung: 10 papers), Inner Ear (6 papers), *Bone Biology
(35 papers), *Stones and Crystals (35 papers);

Several Biomaterials Related Programs organized under "*Cells and
Materials": (1) Skeletal Tissue / Biomaterials; (2) Foreign Body
Reactions; (3) Biointerfacial Reactions at Biomaterial Surfaces;
(4) Innovative Drug Delivery Systems; (5) Blood-Related
Biomaterials; and (6) Dental Biomaterials. (nearly 70 papers)

----------------

For more information, please complete and return the form below to
SMI:

__ I wish to present at the Scanning Microscopy 1996 meeting
(tentative title and summary on a separate sheet), please send a
Letter of Intent form. IT IS VERY LATE NOW!

__ I cannot present, but am likely to attend the 1996 meeting,
please keep me informed and send me: ___ 1996 Registration / Hotel
form.

__ I can neither present nor attend; please add/keep my name on
your mailing list. ___ Send me a mailing list form.

Please send: 1996 program fliers (*list programs here):



___ Instructions for Authors / ___ Major subject index / Table
of Contents for: ___ Scanning Microscopy / ___ Cells and
Materials / ___ Food Structure;

Flier on Scanning Microscopy:

___ Supplement 6, 1992 ("Signal and Image Processing in
Microscopy and Microanalysis");

___ Supplement 7, 1993 ("Physics of Generation and Detection of
Signals Used for Microcharacterization");

___ Supplement 8, 1994 ("Science of Biological Microanalysis")

___ Flier on the special issue: Interface Formation and Dynamics in
Layered Structures (Scanning Microscopy, Vol. 8, no. 4, 1994).

___ Flier on 1996 Pfefferkorn Conference on Electron Image and
Signal Processing, May 18-22, 1996 at Silver Bay, New York


Name

Title

Organization / Company


Address

City/State

Postal Code

Country

Phone

FAX

E.mail

Date




From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Mon, 11 Mar 1996 16:19:46 -0600
Subject: SEM: videotaping from SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130502ad6a56342c43-at-[131.230.97.68]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We have a Hitachi 2460N variable pressure SEM and wish to do videotaping.
The unit has a jack called "Video Out" that we use to obtain quick prints
from a video printer. When we attempted to video tape from this jack, we
obtained a poor quality image (dim, blurry, noisey, etc.). It appears that
this might work but that some intervening device (signal booster) may be
needed. Anyone have any experience with this? Many thanks.

#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: Ginger Baker :      lizard-at-okway.okstate.edu
Date: Thu, 7 Mar 1996 17:02:48 -0600
Subject: OMS Spring Workshop, April 12, 1996, OSU, Stillwater, OK

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0

On Friday, April 12, 1996 the annual Spring Workshop of the Oklahoma
Microscopy Society (OMS) will be held in Room 246 in the Noble
Research Center at Oklahoma State University (OSU), Stillwater,
Oklahoma.



The main focus of the workshop is "Digital Imaging and Image
Processing" and is being presented by Dr. David Bright, 1996 MSA
Speaker and a Research Chemist from the National Institute of
Standards and Technology in Gaithersburg, MD. The workshop will
include some of the following topics:

Image Processing for Microscopists: Practice and Pitfalls
Image Processing for the Macintosh: (free software available)
What are digital images and why would one want to process them?
Introduction to NIH Image and MacLispix
Introduction to Image Processing
-aspect ratio- "correction" for tilt or non square pixels
-background subtraction and normalization
-direction of gradient
-euclidian distance map
-mathematical morphology
-linear hough transform

Other items of business:

(1) Student Best Electron Micrograph Contest: Don't forget to
bring those prize-winning pictures with you to the workshop. Electron
micrographs will be judged by attendees at the workshop. The winner
will receive a $50.00 award and the micrograph will appear on the
cover of the Fall 1996 OMS Newsletter. Second place winner will
receive an award of $20.00.

(2) The Fall 1996 Technical Meeting in conjunction with the
Oklahoma Academy of Science will be discussed. This year it will be
held at the Oklahoma State University-OKC on Friday, November 8, 1996.
Abstracts are due Monday, September 16, 1996.

(3) Final call for the submission of micrographs suitable for the
new OMS t-shirt. Bring your entries to the spring workshop or mail
them to Ginger Baker, Sec. before that date. This contest is not
limited to students. The winner will receive a first edition of this
T-shirt absolutely free. Remember that the T-shirts should be
considered part of OMS' outreach efforts: they will be seen mostly by
the non-microscopist public. In the now-famous words of our OMS
president, "No hexamethylchickenwire lattice structures, please!".

(4) Discussion on the OMS web page expertly compiled and edited by
Dr. Scott Russell, University of Oklahoma.

Address: http://www.uoknor.edu/electron/oms/

From this site, you can travel to a variety of pages for up-to-date
microscopy news and information!

(5) Various meetings and shortcourses coming up during the year.
Also, updates on the gradeschool microscopy kits and their use in the
elementary classrooms.



Preregistration for the OMS Spring Workshop is $5.00 for members,
$10.00 for student non-members, and $15.00 for non-members. After
April 10 add a $5.00 late fee. On site registration begins at 8:30
a.m. Lunch is included in the fee.



Lunch: smoked turkey and brisket with barbecue sauce, cole slaw, baked
beans, relish, wheat/white bread, and peach cobbler.



Parking: Lot 96: Each person driving to the workshop must have a
visitor parking permit to hang in the vehicle with the designated lot
number.

For more information contact either:

Wanda L. Edwards Ginger R. Baker
President-Elect, 95-96 Secretary/Treasurer, 94-96
Email: kmk285-at-vms.ucc.okstate.edu Email: lizard-at-okway.okstate.edu
Phone: (405) 744-7415 Phone: (405) 744-6765
FAX: (405) 744-5275 FAX: (405) 744-8263





From: TKWZ51A-at-PRODIGY.COM (MR ALEXANDER W GINGELL)
Date: Thu, 07 Mar 1996 17:40:13 EST
Subject: TN5500

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Alex W. Gingell * EMC.Ver #2.10P ] --

Thanks to those who replied to my request for used Tracor TN500 EDS
systems for sale.

Unfortunately all my e-mail from last week has been destroyed! Please
re-send.

Best regards,

Alex Gingell
Tha





From: tom von moltke :      tomm-at-iscorltd.co.za
Date: Tue, 12 Mar 1996 15:14:37 +0200
Subject: unsubcribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {314578BD.63E8-at-iscorltd.co.za}





From: jjt-at-vet.purdue.edu (Dr. John Turek)
Date: Tue, 12 Mar 1996 08:09:04 -0500 (EST)
Subject: sectioning titanium alloy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have had a request from a materials science colleague to section a
titanium alloy. He wants a blockface of several square mm and a thickness
of 10 microns. Is this possible to do? Is there anyone out there who would
be willing to do this for a fee? Thanks.

John Turek
Purdue University
Dept. Basic Medical Sciences





From: klivi-at-jhu.edu (Kenneth JT Livi)
Date: Tue, 12 Mar 1996 10:09:29 -0400 (EDT)
Subject: TEM and SEM - Facility User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Arthur Hand wrote:

} I would appreciate any information on current user fees at university EM
} facilities. My institution has requested a comparison of our fees with those
} of other facilities. Thanks for your help.

At The Johns Hopkins Department of Earth & Planetary Sciences, we set the
fee 11 years ago for in-house use at $35/hr for either the TEM or the
microprobe. Add additional fees for operators (cost depending upon whether
operator is student, staff, or faculty). TEM film is included in hourly
rate, but polaroid film is not supplied for EMPA use and holey-C grids are
charged for TEM use. Time starts when filament is turned on. This
includes calibration time for the EMPA.

Cheers,

Kenneth JT Livi
Department of Earth and Planetary Sciences
34th and Charles Streets
The Johns Hopkins University
Baltimore, Maryland 21218
(410) 516-8342 (voice)
(410) 516-7933 (fax)
klivi-at-jhu.edu (e-mail)






From: jlong-at-electrosource.com (James Long)
Date: Mon, 11 Mar 1996 14:12:31 -0600
Subject: Re: photographic processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential
Microscopy ListServer {Microscopy-at-sparc5.microscopy.com}
Message-Id: {960312113205.628-at-cliff.ml.wpafb.af.mil.0}

At 10:12 AM 3/7/96 -0600, you wrote:
} The best processor out there is the Mohrpro. We have been using one
} for about 8 years and recommend it when ever we can. It is easy to
} use and maintain, 2 min. dry to dry, permanent (not stabilized) and
} uses RC paper and Kodak Fixer.

I must respectfully disagree. I was personally responsible for implementing
both an Ilford 2150 and a Mohr Pro 8" Model in a multiuser lab. While I
liked the Mohr and our users were mostly happy with it, from an ease of use
and maint. point of view, I would have to go with the Ilford.

The 2150 ran flawlessly for several years, then needed minor (in-house)
maintenance - float sensors went bad. It will process up to 20" wide sheet,
in 90 secs. dry-dry, and the rollers stay submerged in the chemistry, which
seems to keep them cleaner (no dried on chemicals to deal with).

The Mohr has some rollers partially exposed, which often was a source of
roller marks, and were more difficult to clean. We used it primarily for
negatives (TEM & SEM). It performed well, just not as smoothly as the Ilford.

Overall, I had very good support from both Ilford and Mohr (though neither
produced problems that really tested the waters). I would recommend both,
but for different apps - Ilford for RC Prints (does not do film), and the
Mohr for negs. With the caveat that the Mohr will require more hands-on time.

JCL
=========================================================
James C. Long
Manager/Materials Analysis Lab
Electrosource, Inc.
512-445-6606
jlong-at-electrosource.com





From: Ann-Fook Yang :      YANGA-at-em.agr.ca
Date: Tue, 12 Mar 1996 13:27:46 -0500
Subject: Cryo accessory for Ultracut E microtome

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s1457b18.020-at-EM.AGR.CA}
X-Mailer: Novell GroupWise 4.1

I am looking for a used set of cryo accesssory for Reichert
Ultracut E microtome. If anyone has one and willing to sell,
please contact me directly. Thank you.

Ann Fook Yang
e-mail: Yanga-at-em.agr.ca
Phone: 613-759-1638





From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 12 Mar 1996 13:23:19 -0400
Subject: RE-Heater calibration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1385485336.93602-at-mse.engin.umich.edu}

Subject: Time: 1:09 PM
OFFICE MEMO RE:Heater calibration Date: 3/12/96

We just went through the problem of trying to calibrate a heating device for
TEM specimens. In the process we made use of some very fine thermocouples
which we found can be purchased at a very reasonable price from the Cole
Parmer Company (7425 N. Oak Park Ave., Chicago, IL 60648; 800-323-4340).
These are bare wire thermocouples (which means they are very flexible and
easy to get into awkward places), and they are available in wire diameters
from 0.03" down to 0.005", and in several combinations of wires for different
temp ranges and applications.





From: Walter Arno Mannheimer :      wamann-at-metalmat.ufrj.br
Date: Tue, 12 Mar 1996 15:19:11 EST3EDT
Subject: ISO summary

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopy members,
I have been in correspondence with Jordi Marti. He has been having
problems with email, and has asked me to broadcast that the ISO
summary (which was already posted) did not make it to the system. It
will be reposted as soon as problems are solved (a few days)
Prof.Walter A.Mannheimer
Metallurgy and Materials Engineering
Federal University of Rio de Janeiro
POBox 68505
21945 Rio de Janeiro Brazil
Voice 5521 280-7443 (Dept.office) 5521 590-0579 (direct)
Fax 5521 290-6626 Email: wamann-at-metalmat.ufrj.br




From: Bill Sgammato, NORAN Instruments, Inc.
Date: 12 Mar 96 15:54:41 EST
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Please add me to the mail list for Microscopy.
My E-Mail address is 103506,222-at-compuserve.com

Thank You,

Bill Sgammato





From: John Lafeber :      jlafeber-at-rust.net
Date: Tue, 12 Mar 1996 17:34:44 -0500 (EDT)
Subject: unsubcribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubcribe
***********************************************

John C. Lafeber, P.E.
4165 Stein Road
Ann Arbor, MI 48105

Phone: 313-761-8819
Fax: 313-761-9943
E-mail: jlafeber-at-fe3.rust.net
Web: http://www.rust.net/~jlafeber
FTP (Pick up file location) ftp.netcom.com/pub/jl/jlafeber

"People have vision, computers should too."





From: John Lafeber :      jlafeber-at-rust.net
Date: Tue, 12 Mar 1996 17:34:53 -0500 (EDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe
***********************************************

John C. Lafeber, P.E.
4165 Stein Road
Ann Arbor, MI 48105

Phone: 313-761-8819
Fax: 313-761-9943
E-mail: jlafeber-at-fe3.rust.net
Web: http://www.rust.net/~jlafeber
FTP (Pick up file location) ftp.netcom.com/pub/jl/jlafeber

"People have vision, computers should too."





From: zaluzec-at-Microscopy.Com (Nestor J. Zaluzec)
Date: Wed, 13 Mar 1996 19:16:54 -0600
Subject: Test Message from Nestor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry for the traffice folks, but we had a problem
today on the Server. This is just a test message
so that I can try to locate the source.

Nestor
Your Friendly Neighborhood SysOp






From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Thu, 14 Mar 1996 08:49:45 +0000 (GMT)
Subject: Re: RE-Heater calibration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Wil:
I read your e-mail message about heater calibration with interest. You
might like to seek out a paper some colleagues and I wrote about 20 years ago
Scanning Electron Microscopy 1976 Part I pages 83-90 in which we describe
how we fabricated some very thin film thermocouples for use in the SEM.
The women who did the work was extraordinarily patient and very nimble
with her fingers. They worked, and showed us that a frozen sample kept at
ca. 150K did not heat up when irradiated with the electron beam.

Patrick Echlin
Cambridge




From: James S MArtin :      James.S.Martin-at-williams.edu
Date: Thu, 14 Mar 1996 09:12:12 -0500 (EST)
Subject: polaroid back for sale/trade

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603141323.IAA28742-at-hoh.mbl.edu}

We have an Olympus polaroid camera back that has never been out of the
box. The part numbers, descriptions and 1990 list prices are:

PM-DL-W adapter for large format back 436.00
PM-CP-W 3 1/4 x 4 1/4 in. Polaroid back 260.00

If you are interested in purchasing these items or trading for something
... please contact me directly.

James Martin




From: Jake Schaper :      Jake_Schaper-at-chdqm.sps.mot.com
Date: 14 Mar 1996 09:38:03 -0700
Subject: Subject-SEM Repair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1385325847.24632-at-chdqm.sps.mot.com}

Can anyone refer me to a reliable independent SEM service company that can
repair an S200? I'm in the Phoenix area, so a company near here or the west
coast would be desirable.


**********************************************************
Jake Schaper
ASIC Product Analysis Lab
Motorola, Inc.
1300 N. Alma School Rd. Chandler, Arizona 85224
Mail Drop CH240
Phone 602-814-4756
**********************************************************






From: RNBALDUC-at-ARCRIDE.EDU.AR
Date: Thu, 14 Mar 1996 14:53 -0300
Subject: searching SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear Microscopysts

We are searching an used SEM for received it in donation .The SEM will be
used for Research and Education in our School.
Please, we need know the model, year of fabrication, possibility of work,
and other information available.
We pay all the costs of shipmen.
Please, contact me for any questions.

E-mail : RNBALDUC-at-arcide.edu.ar

Address:

Fernando Balducci
Laboratory of Electron Microscopy
School of Bioengineering.
National University of Entre Rios.
C.C 57 Suc 3
Parana - Entre Rios
Argentina.





From: klivi-at-jhu.edu (Kenneth JT Livi)
Date: Thu, 14 Mar 1996 14:47:28 -0400 (EDT)
Subject: Phantom EDS low-energy tails

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings to all listeners,

I have an interesting problem that I hope someone has some insight on. I
have a light element EDS detector attached to a Philips 420 Supertwin, and
processed pulses are analyzed by the DTSA/4Pi/Mac system. I have observed
an unusual phenomenon that is interfering with our ability to model the
shapes of characteristic peaks and backgrounds. All the peaks exhibit
low-energy tails (LETs) that come and go depending upon a few parameters.
They are:

(1) After a day of the detector bias being on but no X-rays enetering the
crystal, the first spectrum of that day will have large LETs (} 3% of total
peak area) that decrease with further exposure to X-rays.
(2) LETs can be reduced if the bias is turned off for some time, but they
will reappear after a day.
(3) Conditioning the crystal only effects LETs as would turning off the bias.
(4) Previous detectors have not shown as severe LET problems.
(5) The amount of LETs can be reduced by lowering accelerating keV.
(6) Every single component of this system has been replaced, including the
crystal (3 times), but excluding the Mac computer and the TEM. This has
had no effect.
(7) There appears to be no problems related to ground-loops.

Observations 1 through 4 suggest a problem with the current detector
system, but 5 and 6 may imply a problem with the scope alignment. Has
anyone experienced similar problems with their detector? I have heard
rumors that this has appeared on other TEMs, but is not a problem on SEMs
and EMPs.

Many thanks to anyone with ideas.

Kenneth JT Livi
Department of Earth and Planetary Sciences
34th and Charles Streets
The Johns Hopkins University
Baltimore, Maryland 21218
(410) 516-8342 (voice)
(410) 516-7933 (fax)
klivi-at-jhu.edu (e-mail)






From: krogers-at-ecn.purdue.edu (Kirk Rogers)
Date: Thu, 14 Mar 1996 16:24:04 -0500
Subject: WTB: X-Ray detector

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopy Folks-

We would like to request that if someone has an old EDS unit that
will do qualitative analyses and hook up to a PGT detector, we'd be
interested in looking at it and possibly purchasing. We don't need fancy
bells, whistles, image analysis, dot-mapping ect. Just a basic unit that
will take an x-ray spectrum and do some simple analyses.

The names/contact info. of reliable used equipment vendors would also be of
assistance.

Thanks,



-Kirk
_____________________________________________
Kirk Rogers krogers-at-materials.ecn.purdue.edu
OR kirk.a.rogers.1-at-purdue.edu
Purdue University, School of Materials Science and Engineering,
1289 MSEE building, W. Lafayette, IN 47907-1289
OFFICE: 317-494-8751 FAX: 317-494-1204
http://materials.ecn.purdue.edu/~krogers






From: Daniel L. Callahan :      dlc-at-owlnet.rice.edu
Date: Thu, 14 Mar 1996 16:09:13 -0600 (CST)
Subject: Si transmission web-page

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Good Day:

I have placed a picture of a flat-wedge of optically transmitting (100)
silicon on a web page: the image has thickness scale bars from 0 to 10
micrometers. This isn't an ideal sample: for example, there are no
interference fringes visible in the submicron regime. However, it should
serve as a starting reference. Please let me know what you think. The
image is at

http://www.owlnet.rice.edu:80/~dlc/silicon.html

I am considering making a higher angle wedge hoping to preserve more of
the thin region and also taking an image under a Na lamp as advised by
Ron Anderson. Other suggestions are welcome as would other images,
particularly of other orientation.


Daniel L. Callahan
Department of Mechanical Engg. and Materials Science
Rice University
dlc-at-owlnet.rice.edu




From: GANTZ-at-med-biophd.bu.edu
Date: Thu, 14 Mar 1996 18:18:06 -0500 (EST)
Subject: Vitreous Ice Layer Thickness Determination

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Everyone:
I'm working on some projects using vitreous ice cryomicroscopy and
would like to be able to determine the thickness of final ice layers on
holey films. In general, I've heard of a couple of methods: 1) Burn a
small hole, tilt, photograph; or 2) relative density by densitometry of
negatives. I would be appreciative to receive details and references
on these methods or other methods being used. What sort of accuracy
should one expect from the various methods? Thanks for your help.
Don Gantz
Boston Univ. Med School
e-Mail: gantz-at-med-biophd.bu.edu




From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Thu, 14 Mar 1996 19:30:55 -0600
Subject: SEM Videotaping Problem Solved

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130505ad6e77c4714a-at-[131.230.97.68]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Thanks to all who sent helpful suggestions regarding our inability to
videotape from our S2460N variable pressure SEM. The problem was caused by
a bad video input on our new video recorder (same brand as the microscope,
oddly enough). When we used a different recorder, things worked
beautifully. We're smilin' again.

#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Fri, 15 Mar 1996 13:54:35 +1200
Subject: Creutzfeld-Jakob disease biohazards

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listserver readers,
We are a university EM unit primarily concerned with teaching and
supporting research across many disciplines. The local hospital have a
technician based in our Unit who processes their diagnostic cases for EM.
Recently it came to my attention that he had processed a brain biopsy from
a suspected Creutzfeldt-Jakob disease patient. What really concerned me was
that :
1) We had not been informed that the sample was in the lab
2) The hospital technician did not appear to be aware of the high-risk
nature of the disease.

This incident has highlighted two areas of concern in our Unit. Firstly
we have no mechanism in place that requires the hospital to inform us when
high risk biohazards come through our lab. Secondly, our mechanisms for
dealing with "high risk" samples and disposing of such material need to be
re-evaluated and improved.

My questions are:

How do other labs that are have high risk samples going through them
monitor the risk of the samples coming into the lab?

How do other labs handle and dispose of high risk biohazards?

Allan Mitchell
Richard Easingwood



Richard Easingwood
South Campus Electron Microscope Unit
School of Medical Sciences
University of Otago
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

SOUTHERNMOST E.M UNIT IN THE WORLD















From: NICK SCHRYVERS :      nick_schryvers-at-ruca.ua.ac.be
Date: 15 Mar 1996 08:57:19 +0100
Subject: Re: TEM on archeol. etc.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1385241999.36229-at-ematserv.ruca.ua.ac.be}
"microscopy-at-Sparc5.Microscopy.Co" {microscopy-at-Sparc5.Microscopy.Com}
X-Mailer: Mail*Link SMTP-QM 3.0.1

Reply to: RE} TEM on archeol. etc.

Dear Amy and others,
Good question. I have been looking (via the Current Contents for Physics and
Chemistry) for TEM studies on historical art and archeological material for
more than five years now and have never found any decent reference. It's
always SEM. Of course the fact that for TEM you'd have to destroy part of the
object is a serious problem. Nevertheless, as small quantities can be
sufficient I agree that it should be considered.
Could you please inform the group of the reactions that you get (maybe by
posting a compilation after a few weeks).
Thanks a lot.
Nick Schryvers








From: Katherine.S.Connolly-at-Dartmouth.EDU (Katherine S. Connolly)
Date: 15 Mar 96 08:36:53 EST
Subject: high-risk samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {26557033-at-dancer.Dartmouth.EDU}

In this era of HIV and other potentially lethal contents of samples all of ours
are treated as high-risk.
Kate Connolly




From: Richard Lee :      richard_lee-at-QMGATE.ANL.GOV
Date: 15 Mar 1996 08:50:46 -0600
Subject: SMSI March meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1385242375.2626-at-qmgate.anl.gov}
"Ankur Purohit" {ankur-at-anl.gov}
X-Mailer: Mail*Link SMTP-QM 3.0.2

Thanks Jonathan for agreeing to be the invited speaker on March 29. Copies of
the notice are on their way to you. The title,"Extremophiles - organisms at
the extremes" is intriguing. By the way can you explain for us what the CMB is
and does?
If you need info on how to get to McCrone Research Institute, I will send it
to you.

8:38 AM




From: Dave King (607)857-1248 T37/257-3 Ext DEKING-at-VNET.IBM.COM
Date: 15 Mar 1996 12:47:20 EST
Subject: 10 Micron Ti Section

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: Microscopy ListSer {Microscopy-at-MSA.Microscopy.Com}
cc: BLAZEY --ENDVM1 Blazey, W.C. EMMI --ENDVM5 F. Emmi
WAITKUS --ENDVM1 Waitkus, C. CONROW --ENDVM5 K.M. (Karen) Conro


We're proud of our capabilities, but near 10 micron thickness
controlled over several mm is beyond our metallographic
sectioning capability and our microtome, both! We could work with
you on an experimental basis, but your goal is way out there.

{
} { {
} } ===========} Dave King { { {
} } } ------------------------------------------------------ { { { {




From: geosclmr-at-showme.missouri.edu (Lou Ross)
Date: Fri, 15 Mar 1996 12:01:45 -0600
Subject: X-sectioning of filter paper.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to all who responded.

Lou Ross

101 Geological Sciences Bldg.
University of Missouri
Columbia, MO 65211
(314) 882-4777, 882=5458 fax

'First came fire/Then came light/
Then came feeling/Then came sight'






From: STEELE-at-KRDC.INT.ALCAN.CA
Date: Fri, 15 Mar 1996 08:57:23 -0500 (EST)
Subject: Prep of ceramic fibres (TEM)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



--Boundary (ID g1sXwhSkJTbyydWcZFNosA)
Content-type: TEXT/PLAIN
Content-transfer-encoding: QUOTED-PRINTABLE

I'm guesing by Nestor's test message that this didn't get through the=
first=20
time, so I'll try again. Apologies to any who are receiving this for =
a second=20
time.

I received a few requests for details on my electroless Ni route for=
=20
fibre/powder preparation for the TEM. So here, culled from the lit., =
is the=20
method I've settled on:

---------------------------------------------------------------------=
--------=20

I'll start with the technique I use for electroless coating, then giv=
e a few=20
references for electroplating, and then the method I use. For ceramic=
powders I=20
I recommend electroless followed by electroplating.

Electroless Ni coating:

Apparatus: Hot plate / stirrer
3 stirring bars
Funnel, stand, Whatman 40 filter papers
4 250 ml beakers (glass) (Xtremely clean!!!)

3 solutions, A,B,C.

Solution A: Add 0.1 gm SnCl2 to 100ml water pH adjusted to 2 with=
HCl.
(sensitising) Use at rm temp.

Solution B: Dilute 2mls of a 5% PdCl2 solution to 100 ml.
(activator) Use at rm temp.

Solution C: Commercial Eleectroless solution made by HARSHAW. Ava=
ilable=20
through Engelhard technologies (416) 821 - 3325.
Comes in 3 parts Alpha 103 A-MB, Alpha 103 A-2, and=
=20
Alpha 103 A-1 replenisher. =20

Make up 100 mls by adding 20 ml A-MB, and 7.5 ml A-2,=
and=20
bringing up to 100 ml with D/I H2O. The replenisher =
you will
need if you use up all of the Ni, follow the destruct=
ions=20
given by Harshaw for gauging this, and replenishing.

This solution is used at 90 - 95=F8C.

Proc: Add aprox 1cc powder to solution A with stirring. Af=
ter=20
2 min. filter and rinse the cake with D/I H2O. Place=
the=20
filter cake and paper in solution B with stirring. A=
fter
2 min filter the sol and rinse the cake with D/I H2O.

Prepare a slurry by scraping the cake from the paper =
and=20
adding no more than 10ml H2O. Pour this into solutio=
n C,
with stirring, allow to plate for aprox. 1 min, then =
filter=20
and wash the powder.

And there you go.

Next, you incorporate the powder into a Eplated Ni sheet in your favo=
rite
plating bath (I use 80 ml Ni Sulphamate + 120 ml D/I H2O + 10 gm Bori=
c Acid +
1gm Nickel Chloride, + 5 drops Hydrogen Peroxide).=20

I use a stainless steel beaker, an "L" shaped cathode to drop the pod=
er on,=20
a high purity Ni anode, heat the sol to 50=F8C, and plate at 200 -300=
mA (2.5V)
DC. You need to plate a thin film(20 min), add just enough powder to=
=20
get an even dispersion, and then let it plate away for 3 to 4 hrs. T=
his=20
should result in a film somewhat in excess of 200=E6m.

A few key references:

S.D. Kirchoff and J. Y. Adkins, Metallography, 20 (1987) 75-87.

M.M. Morra, J. M. Morra, and R.R. Biederman, Materials SCiaence and E=
ngineering,
A124 (1990) 55-64

R. Baumann and M. J. Couper, Prakt. Met. 23 (1986) 140 - 148

R. D. Field and H. L. Fraser, Met. Trans. 9A, (1978) 131 - 134.

---------------------------------------------------------------------=
--------

###############################################################
# #
# Don Steele STEELE-at-KRDC.INT.ALCAN.CA #
# ALCAN INTERNATIONAL #
# Kingston Research and Development Center #
# P. O. Box 8400 #
# Kingston, Ontario Canada K7L 5L9 #
# #
###############################################################



--Boundary (ID g1sXwhSkJTbyydWcZFNosA)--




From: William Tivol :      tivol-at-wadsworth.org
Date: Fri, 15 Mar 1996 11:25:09 -0500 (EST)
Subject: Re: Creutzfeld-Jakob disease biohazards

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Recently it came to my attention that he had processed a brain biopsy from
} a suspected Creutzfeldt-Jakob disease patient. What really concerned me was
} that :
} 1) We had not been informed that the sample was in the lab
} 2) The hospital technician did not appear to be aware of the high-risk
} nature of the disease.
}
} My questions are:
}
} How do other labs that are have high risk samples going through them
} monitor the risk of the samples coming into the lab?
}
} How do other labs handle and dispose of high risk biohazards?
}
Dear Allan & Richard,
Working at a state health department gives us a leg up on such
things. There are procedures in place for proper notification, handling
& disposal. We know what samples are being brought in for examination,
and since it is usually a physician who brings in any potentially danger-
ous specimens, we are made aware of any potential hazards. Usually, the
specimen has already been fixed, stained and embedded, which eliminates
the hazard. There are containers all around our lab for the disposal of
biohazards, and the safety office is responsible for their proper dispo-
sal. Having the hospital technician give you a list of specimens and
their condition (tissue, blocks, sections, etc.) before bringing them in
would be an obvious step. I can think of rationalizations for not doing
this, but maybe you can insist. If you have a safety office, by all
means get them involved. Good luck.
Yours,
Bill Tivol




From: Larry Ackerman :      mishot-at-itsa.ucsf.edu
Date: Fri, 15 Mar 1996 13:55:59 -0800 (PST)
Subject: Re: Video Printer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For decent video encoders (translators) at resonable cost ($200-$400 US) try
Harmonic Research in Paramus, New Jersey at (201) 652-3277.




From: ScottE57-at-aol.com
Date: Fri, 15 Mar 1996 21:04:18 -0500
Subject: Re: Video Printer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603152153.NAA26418-at-holonet.net}

The new Sony printers UP5600 will take PAL or NTSC signals - we handle a full
line - also you could use a frame grabber from a variety of manufactures that
take PAL in and let you output NTSC and use the existing printer. We carry
all these products, please call if you require further assistance.

Scott E. Berman
Advanced Imaging Concepts
Princeton, NJ
Phone(908) 274-1877 x26
Fax(908) 274-1974




From: EvexAnalyt-at-aol.com
Date: Sat, 16 Mar 1996 08:17:22 -0500
Subject: ISO Summary for Allied Signal Jordi Marti

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1385200787.59973-at-QuickMail.Yale.edu}

For those of you waiting for the ISO summary, Jordi Marti (at Allied Signal)
has been having problems with his E-Mail (he can't send mail out) for the
past two weeks. The summary will be mailed as soon as the E-Mail problems
are solved.

The above message was posted by Evex Analytical
as a courtesy for Jordi Marti, Allied Signal




From: James Patrick :      JPATRICK-at-OPUS.MCO.EDU
Date: Sat, 16 Mar 1996 16:30:47 -0500 (EST)
Subject: ISO Summary for Allied Signal Jordi Marti

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: alan-at-macro.mse.uiuc.edu (Alan E. Lucero)
Date: Sun, 17 Mar 1996 08:38:48 -0600
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



unsubscribe




From: Rune Sundset :      runes-at-fagmed.uit.no
Date: Sun, 17 Mar 1996 19:10:39 -0200
Subject: intracellular calcium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603171807.TAA07477-at-petrus.fagmed.uit.no}
X-Sender: runes-at-pernille
X-Mailer: Windows Eudora Light Version 1.5.2
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I would like to know if there is a protocol out there for staining and
measuring intracellular calcium by using EM.

Best wishes,
Rune

======================================================================
Rune Sundset

Dept. of Medical Physiology
Inst. of Medical Biology
University of Tromsoe
N-9037 Tromsoe
Phone : +47 77 67 45 48 or +47 77 64 46 96 Fax : +47 77 64 54 40

----------------------------------------------------------------------
Private
Orneveien 12/213
N-9015 Tromso
Norway
Phone: +47 77 67 45 48
======================================================================





From: HENRY P ADAMS :      hadams-at-nmsu.edu
Date: Sun, 17 Mar 1996 12:05:37 -0700 (MST)
Subject: Fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fellow microscopists, I have to fix some algae (Cyanidium sp.) that are
growing at pH 1.8 for TEM with and without Ruthenium red, to stain
extracellular material. From what I have read glutaraldehyde fixes poorly
at low pH's and conventional fixative buffers do not work below pH 4 (P04).
Also,they can not tell me what the growth media is.
I thought I could add straight glutaraldehyde and then bring up the pH slowly
with NaOH or just rinse and fix normally under the assumption that the
ultrastructure will not be altered, initially, by a higher pH. Does anyone
have any experience with similar organisms or know
of any references that may guide me? I guess I am inclined to try several
approaches, but as is often the case the person has very limited funds
and I was hoping to use someone elses experience (experiments) to reduce
time and labor.
Thanks in advanced,
Hank Adams
EML, NMSU
Las Cruces, NM
hadams-at-nmsu.edu




From: slocombe-at-wsunix.wsu.edu (Peter Slocombe)
Date: Sun, 17 Mar 1996 19:13:40 -0800
Subject: Help! Leica or Zeiss?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Perhaps you can help me with a personal dilemma; I'm about to purchase a
microscope and I've boiled
the choice down to either a Leica Galen III or a Carl Zeiss Standard 20.
The Leica has more bells and
whistles than the Zeiss, and for the setup I'm thinking of it's also about
$1000 cheaper than the Zeiss. I
have not used either and am 500 miles from the closest demo unit. In your
personal opinion would I be
better off, in the long-run, going with the Zeiss? Is one more comfortable
to use than the other in terms
of fatigue? How would YOU compare these two microscopes generally?

Peter Slocombe
Washington State University
slocombe-at-wsunix.wsu.edu





From: Stephan Helfer :      S.Helfer-at-rbge.org.uk
Date: Mon, 18 Mar 1996 09:47:38 BST
Subject: Re: Help! Leica or Zeiss?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Peter
Here at RBGE we have both Leica and Zeiss models, although we don't
have the Galen III. Traditionally we always had Leica models but when
they changed to infinity corrected optics, traditions didn't help any
more, and so we started looking afresh with an unbiased (obviously
infinity corrected) view. Our choice was Zeiss, partly because of
better optics although some of us find the Zeiss microscopes slightly
less user friendly, and I believe the Leica optics have improved now.
Here is my suggestion: You may live 500 miles away from a demo unit,
but why don't you ask both Leica and Zeiss to leave the microscope
with you for a week or so; this is the only way you can really judge
for yourself which one you would prefer for your own specimens. The
physics of microscopy are pretty well sussed out these days, so what
you want to pay extra for is durability (less trimmings normally
means less trouble) and ease of use (more trimmings sometimes help).

Cheers

Stephan

Yours sincerely

Dr Stephan Helfer, SSO
Mycologist / Plant Pathologist

Royal Botanic Garden Edinburgh, Inverleith Row, EDINBURGH EH3 5LR,
Scotland UK

http://www.rbge.org.uk

phone: +44 (0)131 552 7171 ext 280
or +44 (0)131 459 0446-280 (direct digital VoiceMail line)
fax: +44 (0)131 552 0382
============================
1896
* BRITISH
* MYCOLOGICAL
* SOCIETY
1996
A century of fungal science
============================




From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 18 Mar 1996 10:04:59 EST
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Has anyone heard of the slow viruses that can survive gluteraldehyde fixation?

The infectious agents that are able to survive fixation are the
prions. These agents contain protein only...no nucleic acid. Among
these are Creutzfeld-Jakob, scrapie, kuru,
Gersten-Straussler-Schenker, BSE (mad cow disease) and apparently
some forms of Alzheimers.

It is known that formaldehyde does not inactivate them, not surprising
since formaldehyde is one of the least cross-linking of all
fixatives, so far as proteins are concerned. My understanding is
that glut. does inactivate them, but don't take my word for it.

} On a similar note, there was an MSA-sponsored speaker a couple of years ago who
} was encouraging EM labs to make money by offering virus identification sevices
} to medical centers. The preparation protocol included advice to accept unfixed
} (ie. infectious), unidentified material and prepare negative stained samples on
} the bench in the laboratory.

We are a CLIA and State certified human clinical laboratory, licensed for
negative-stain virus identification. This has always been part of
our function as a veterinary EM lab, and for about 5 years we have
offered it as a service to the community. We do make a considerable
amount of money from it. We accept samples only from certain
reference laboratories and practitioners and these are only stool
samples, mostly (} 95%) from infantile diarrheas. Our submission form has an
entry for any unusual precautions that must be taken (ie HIV,
hepatitis, etc). Our understanding with our clients is that we don't
process such samples.
W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: gwisler-at-asrr.arsusda.gov
Date: Mon, 18 Mar 1996 12:38:43 -0500 (EST)
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Title: Biological Science Technician (Plants)
Lab: USDA-ARS, Salinas, CA

The USDA-ARS is seeking a biological science technician (plants)
(GS-404-7, 8, or 9) for the Crop Improvement and Protection Research Unit
in Salinas, CA. The incumbent will share responsibilities in electron
microscopy of plant virus infections for ultrastructural characteristics.
The incumbent will also assist in research involving molecular, serological,
and biological studies of several plant viruses infecting sugarbeet and
vegetables. Candidate must have a knowledge of electron microscopy,
plant virology, and knowledge of microbiological techniques. Must be a U.S.
citizen. Bachelors degree is desirable. Salary is commensurate with
experience ($24,610-39,140 per annum).
For information regarding research program contact Gail C. Wisler or
James E. Duffus (408)755-2835. For information regarding application
procedures/forms contact Tom Nelson (408)755-2810. Applications must be
postmarked by May 6, 1996. The USDA is an equal opportunity employer.





From: azriel gorski :      azrielg-at-cc.huji.ac.il
Date: Mon, 18 Mar 1996 20:30:11 +0300 (GMT+0300)
Subject: Re: Help! Leica or Zeiss?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I find it hard to believe that no such scopes have been *sold* closer
than 500 miles away. Ask your dealers, if they have sold such scopes, and
if they will contact those people and see if they will be willing to have
you inspect the scope.

Both companies make *excellent* products. At this quality, what feels
right for *you* is very important. So is the *service* you get from your
agent. Also check what scopes your facility has. The ability to "borrow"
a specific "widget" down the hall should be weighed in the equation.

As another user suggested, the agent may be able to lend you a scope to try.

All these conversations should also give you a pretty good idea of how well
you can work with the agent after the sale.

Good luck

Azriel Gorski, Head
Optical Microscopy Laboratory
Division of Identification and Forensic Science
Israel Police




From: Amy G Aslamkhan :      aslamkha-at-hawaii.edu
Date: Mon, 18 Mar 1996 10:22:34 -1000
Subject: TEM-Microwave fixation of marine organisms

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a protocol for microwave fixation of marine organisms?
I am considering this technique for fixing some crustacean tissue
(hepatopancreas), and a colleague of mine is considering microwave fixing
marine algae.

Amy Aslamkhan
University of Hawaii
aslamkha-at-hawaii.edu




From: biln-at-ozemail.com.au (bill neill)
Date: Mon, 18 Mar 1996 20:13:01 +1000 (EST)
Subject: em listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'd like to subscribe to the em listserver, please

Bill Neill
Leica Instruments





From: J-CRAIG-at-csu.edu (Joyce Craig)
Date: Mon, 18 Mar 1996 16:05:17 -0600 (CST)
Subject: bio-physics and microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are attempting to start a bio-physics program at Chicago State University. We need feed back as to the role of physics in biological sciences problem solving. What is the trend? Is it accurate to say that biology and physics are becoming intertwined
? I personally see a need for a greater understanding of physics when working with the newer technology in electron microscopy in biological science but what do you think?





From: MSC18 :      MSC18.MSC18-LARS-at-bbsrc.ac.uk
Date: Mon, 18 Mar 1996 21:48:05 +0000
Subject: TEM: antibodies to anthocyanidins

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Via: uk.ac.bbsrc; Mon, 18 Mar 1996 21:48:50 +0000
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Mon, 18 Mar 1996 21:50:50 +0000
X400-Received: by mta arcs.bbsrc.ac.uk in /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Mon, 18 Mar 1996 21:50:50 +0000
X400-Received: by mta LARS.MUAS in /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Mon, 18 Mar 1996 21:46:48 +0000
X400-Received: by mta LARS in /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Mon, 18 Mar 1996 21:47:34 +0000
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; converted (ia5-text); Relayed;
Mon, 18 Mar 1996 21:48:05 +0000
Return Requested)







From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Mon, 18 Mar 1996 16:40:07 EST
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} As far as I know Alzheimers disease is not infectous---- Where did you get your
} information about Alzheimers disease?

This information is at least 4 years old and may not be correct.
About 4 years ago (maybe a little longer) there was a lengthy review
article in either New England Journal of Medicine or JAMA regarding
the slow human neurological disorders. At that time, there were at
least 5 recognized diseases that were classified as "Alzheimer" or
"Alzheimer-like". One of these was suspected to be of prion origin,
although by now it may be classified as something else.

One thing to consider if you are involved in studies involving human
(or any mammalian) brain tissue is that in years past, prion diseases
were commonly misdiagnosed as Alzheimers, and as humans (even
pathologists) are not perfect, this still may be the case.



W. L. Steffens, Ph.D
Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: jean ross :      jeanross-at-emiris.iaf.uiowa.edu
Date: Mon, 18 Mar 1996 12:55:06 -0600 (CST)
Subject: sectioning teeth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have been working on a project with a dental grad student involving
ultramicrotomy of undecalcified human teeth. The samples were prepared by
slicing them into approx. 100um slices and then embedding them in Epon. They
section just great but something happens when they're picked up on grids which
causes just the dentin portion of the section to wrinkle. I use formvar coated
200 mesh copper grids, stabilized with carbon and glow discharged. I spread the
sections out in the boat with chloroform (I have used xylene too) and have
picked them up every differnt way I could think of. Is it my support film?
Would a thicker section be desirable (sections are usually 90-95 nm)? It is
only the dentin part of the section that has this problem, the Epon and
composite resin reconstruction are perfectly flat. Does anybody have any
suggestions?

Thanks in advance,

Jean Ross
Central Microscopy Research Facility
University of Iowa
85 EMRB
Iowa City, IA 52242
(319) 335-8142
Web site: http://www.uiowa.edu/~cemrf







From: SMithP-at-agresearch.cri.nz
Date: Tue, 19 Mar 1996 17:36 +1200 (NZST)
Subject: Fluorescent in situs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




We are interested in trying some fluorescent in situ hybridisation on tissue
sections. We would be using fetal sheep tissue and would be probing with
riboprobes to growth factors and hormone receptors mainly.These probes
generally work well (as far as any in situ does) when radioactively
labelled, but we would be keen to get away from radioactive labels and would
really like to try multiple labelling. It seems very difficult to find
references describing the technique applied to tissue sections so any help,
advice ,sources of information or contacts would be much appreciated.

Thanks for your time

Peter Smith
AgResearch Wallaceville
PO Box 40063
Upper Hutt
New Zealand.

E-mail smithp-at-agresearch.cri.nz




From: Rune Sundset :      runes-at-fagmed.uit.no
Date: Tue, 19 Mar 1996 09:47:34 -0200
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603190844.JAA22043-at-petrus.fagmed.uit.no}
X-Sender: runes-at-pernille
X-Mailer: Windows Eudora Light Version 1.5.2
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

subscribe

======================================================================
Rune Sundset

Dept. of Medical Physiology
Inst. of Medical Biology
University of Tromsoe
N-9037 Tromsoe
Phone : +47 77 67 45 48 or +47 77 64 46 96 Fax : +47 77 64 54 40

----------------------------------------------------------------------
Private
Orneveien 12/213
N-9015 Tromso
Norway
Phone: +47 77 67 45 48
======================================================================





From: SveEn-at-pai.liu.se (Sverker Enestrom)
Date: Tue, 19 Mar 1996 09:13:29 +0100
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} This information is at least 4 years old and may not be correct.
} About 4 years ago (maybe a little longer) there was a lengthy review
} article in either New England Journal of Medicine or JAMA regarding
} the slow human neurological disorders. At that time, there were at
} least 5 recognized diseases that were classified as "Alzheimer" or
} "Alzheimer-like". One of these was suspected to be of prion origin,
} although by now it may be classified as something else.
}
} One thing to consider if you are involved in studies involving human
} (or any mammalian) brain tissue is that in years past, prion diseases
} were commonly misdiagnosed as Alzheimers, and as humans (even
} pathologists) are not perfect, this still may be the case.

AD and prion diseases share the presence of amyloid precursor proteins
(beta-PP, APrP) but they are distinct diseases, both starting at the synapse
(beta-PP and PrP are proteins of the neuromuscular junction and CNS synapse).
In some of the prion diseases prion amyloid plaques are seen together with
paired helical filaments (NFT) making them "Alzheimer-like" as Budy writes.


*********************************************************
Sverker Enestrom M.D., Ph.D.
Department of Pathology
University of Linkoping, Sweden
Phone: +46 13 22 15 20
Fax: +46 13 13 22 57
*********************************************************






From: Reinhard Windoffer :      reinhard.windoffer-at-zoologie.uni-hamburg.de
Date: Tue, 19 Mar 1996 13:59:52 -0800
Subject: help: subscription to microscopy listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {314F2E58.147A-at-zoologie.uni-hamburg.de}

Can you please send me the details for subscription to the microscopy
news group.


Reinhard Windoffer




From: ENERGY BEAM SCIENCES, INC :      75767.640-at-CompuServe.COM
Date: 19 Mar 96 08:13:57 EST
Subject: Re: TEM-Microwave fixation of marine organisms

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Amy-
I don't know of any specific references to microwave fixation of crustacean
tissue, but you might want to experiment with the method developed by Bill &
Ruth Dewel (Appalachian State University) for fixation of tardigrades
(Echiniscus viridissimus). Please contact me by e-mail for the methodology.
General information on microwave fixation for TEM can be found at our WWW site
(http://www.ebsciences.com/).
Steven Slap, Vice-President
75767,640-at-compuserve.com





From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Tue, 19 Mar 1996 08:23:15 -0600
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d02ad74722277f6-at-[128.206.15.200]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Greetings,
My understanding was that you got prion diseases through eating
contaminated samples, hopefully uncommon in the world's em prep rooms. Is
there an airborne transmission route?
Thanks,
Tobias Baskin

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
___ ____ ^ ____ _____ Tobias I. Baskin
/ \ / / \ / \ / University of Missouri
/ | / / \ / / Biological Sciences
/___ / /__ /_____\ / /__ 109 Tucker Hall
/ / / \ ( / Columbia, MO 65211 USA
/ / / \ \ / voice: 573-882-0173
/ /____ / \ \____/ /_____ fax: 573-882-0123






From: MSC18 :      MSC18.MSC18-LARS-at-bbsrc.ac.uk
Date: Tue, 19 Mar 1996 14:47:20 +0000
Subject: TEM sorghum phytoalexins

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Via: uk.ac.bbsrc; Tue, 19 Mar 1996 14:53:40 +0000
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Tue, 19 Mar 1996 14:49:57 +0000
X400-Received: by mta arcs.bbsrc.ac.uk in /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Tue, 19 Mar 1996 14:49:57 +0000
X400-Received: by mta LARS.MUAS in /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Tue, 19 Mar 1996 14:47:06 +0000
X400-Received: by mta LARS in /PRMD=UK.AC/ADMD= /C=GB/; Relayed;
Tue, 19 Mar 1996 14:47:09 +0000
X400-Received: by /PRMD=UK.AC/ADMD= /C=GB/; converted (ia5-text); Relayed;
Tue, 19 Mar 1996 14:47:20 +0000
Return Requested)

Hello,
I am currently doing TEM on sorghum leaves looking at the infection
process of Colletotrichum graminicola. Sorghum leaves produce
anthocyanidin phytoalexins in response to infection and I would like to
locate these compounds under the TEM. I was wondering if anybody out
there had or knew where I could get hold of antibodies to anthocyanidins,
specifically apigeninidin and luteolinidin, or if anyone knew of any stains
for EM which were specific to anthocyanidins.

Phill Wharton

*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*
Phillip Wharton
IACR-Long Ashton Research Station
Long Ashton
Bristol
BS18 9AH
*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*




From: William Tivol :      tivol-at-wadsworth.org
Date: Tue, 19 Mar 1996 12:53:43 -0500 (EST)
Subject: Re: bio-physics and microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} We are attempting to start a bio-physics program at Chicago State
} University. We need feed back as to the role of physics in biological
} sciences problem solving. What is the trend? Is it accurate to say
} that biology and physics are becoming intertwined? I personally see a
} need for a greater understanding of physics when working with the newer
} technology in electron microscopy in biological science but what do you
} think?

Dear Joyce,
Physics has several roles in bio problem solving. There is the
application of physical measurements (ultracentrifugation, calorimetry,
diffraction to name just a few), the use of physics in understanding what
happens to the specimen (radiation damage in EM, physical preparation
techniques like microwaving & sectioning), and, perhaps most important,
the basic viewpoint of physics can aid thinking about bio phenomena--
this has proved to be especially valuable to me. I found that my training
in physics gave me the quantum-mechanical viewpoint from which to view
enzyme-substrate interactions and molecular structure--which gave me in-
sights that my colleagues not trained in physics did not have.
I believe the trend in research is toward interdisciplinary studies,
so physics, among other disciplines, is being applied more and more to other
fields, including biology. I would not single out the bio-phys linkage as
unique (or even one-way). I think that all disciplines are becomming more
fluid with the boundaries getting fuzzier rather than that biology and
physics specifically becomming intertwined.
Understanding the principles of the instrumentation is always a
plus, and the newer technology certainly relies on physical principles.
On the other hand, the newer instrumentation is also becomming more user-
friendly and computer-oriented. This means that the instruments can be
operated merely by clicking on a few menus, and repair of a faulty in-
strument usually means replacing a computer chip rather than disassembling,
analysing which part is broken, making a new part and reassembling. At the
latter level, less understanding of physics may be required to perform the
experiment, but more required to understand it.
Yours,
Bill Tivol





From: Gary Login :      glogin-at-bih.harvard.edu
Date: Tue, 19 Mar 1996 13:45:40 -0500
Subject: TEM-Microwave fixation of marine organisms

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In message {Pine.SV4.3.91.960318100758.25965A-100000-at-uhunix5} Amy G Aslamkhan
writes:
} Does anyone have a protocol for microwave fixation of marine organisms?
} I am considering this technique for fixing some crustacean tissue
} (hepatopancreas), and a colleague of mine is considering microwave fixing
} marine algae.
}
} Amy Aslamkhan
} University of Hawaii
} aslamkha-at-hawaii.edu


Microwave fixation is an ideal method to preserve the soft tissue within a hard
(calcified shell). References appropriate to your application follow:

Berg, C. J., and N. L. Adams. Microwave fixation of marine invertebrates. J Exp
Mar Biol Ecol 74: 195-199, 1984.

Login GR, Dvorak AM. Methods of microwave fixation for microscopy. A review of
research and clinical applications: 1970-1992. Prog Histochem Cytochem
1994;27/4: 1-127.


If you need further assistance or methodological details please do not hesitate
to contact me by calling or sending a fax to the number at the end of this
E-mail message.

Gary Login



Gary R. Login, D.M.D., D.M.Sc.
Assistant Professor of Oral Pathology
Beth Israel Hospital
Department of Pathology
330 Brookline Avenue
Boston, Massachusetts, 02215

glogin-at- bih.harvard.edu
Telephone: 617-667-2034
Fax: 617-667-8676





From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Tue, 19 Mar 1996 14:05:57 EST
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} My understanding was that you got prion diseases through eating
} contaminated samples, hopefully uncommon in the world's em prep rooms. Is
} there an airborne transmission route?

The review article that I described earlier notes some cases of prion
disease in neurosurgeons who had performed procedures on known
infected patients in years past.
Not exactly iron-clad evidence of alternative transmissibility....but
still.

This thread had gotten my interest on the subject piqued again. I'll
see if I can dig out the original article.




W. L. Steffens, Ph.D
Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: lucina-at-xk.com (Lucina Mastro)
Date: Tue, 19 Mar 1996 12:37:33 -0800
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: lucina-at-pop.sirius.com
Message-Id: {v01530500ad74b91082cc-at-[205.134.227.50]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

subscribe






From: kna101-at-utdallas.edu
Date: Tue, 19 Mar 1996 14:04:58 -0600 (CST)
Subject: Re: sectioning teeth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jean:

I've come across a similar problem with epon embedded inner ear
(cochlea). In my case, the soft tissue I need to look at is completely
surrounded by decalcified bone. Several things I've tried seem to help
with this problem: trim away the bone from the surface of the block
before sectioning or increase infitration times, using a harder epon
mix to try and reduce the difference in hardness that occurs in the
block between pure epon, bone and soft tissue. I hope this helps.
P.S. I've had this happen even on bare grids.

Karen Pawlowski
Technician, UT Southwestern Medical Center
PhD student, UT Dallas




From: yimei-at-befvax.uchicago.edu
Date: Tue, 19 Mar 1996 16:47:42 EST
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe Yimei-at-befvax.uchicago.edu




From: Paul Webster :      paul.webster-at-yale.edu
Date: 19 Mar 1996 18:20:46 -0500
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1384862484.65115-at-QuickMail.Yale.edu}

} My understanding was that you got prion diseases through eating
} contaminated samples, hopefully uncommon in the world's em prep rooms. Is
} there an airborne transmission route?

The most common form of transmission for toxoplasma gondii is from eating cat
feces but I am sure no-one would admit to eating that!



Paul Webster, Ph.D.
Center for Cell Imaging
Yale University School of Medicine





From: Jill Craig :      jcraig-at-unbc.edu
Date: Tue, 19 Mar 1996 15:21:46 -0800 (PST)
Subject: unions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi,

The University here has recently decided to unionize the staff. I was
wondering if SEM, NMR, MS technicians are commonly unionized in the
experiences of the group. We were hired into our designated positions
and there really isn't a ladder to climb to higher positions or
designations. Overall, I'm having a difficult time seeing the advantage
of being part of a union although I'm fully aware of the price deducted
from my salary. Is this a common practice in Canada or elsewhere?

Reply to me and I'll summarize.

Thanks,

Jill




From: Colin MacRae :      C.Macrae-at-minerals.csiro.au
Date: Wed, 20 Mar 1996 10:52:25 +1000
Subject: Angle measurement of single crystal faces

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603191826.NAA12061-at-hoh.mbl.edu}

I am using an In_lens Field Emission SEM to image various single crystals.
Can anyone suggest a technique for measuring the angles between single
crystal faces? (Since the IFESEM is an in-lens microscope use of a double
detector system is impossible)

Is it possible to take secondary electron stereo pair images and measure the
true crystal face angles?

Colin MacRae
Scanning Probe Microscopy and Electron Microscopy Section

Commonwealth Scientific and Industrial
Research Organisation. _--_|\ cmac-at-minerals.csiro.au
Division of Minerals / \ +61 3 9647 0296
PO Box 124, Port Melbourne 3207 \_.--._/ fax 61 3 9646 3223
AUSTRALIA

EM units WWW site http://www.minerals.csiro.au/em-unit/





From: Scott D. Whittaker :      sdw-at-biotech.ufl.edu
Date: Tue, 19 Mar 1996 22:49:02 GMT
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 06:20 PM 3/19/96 -0500, you wrote:
} } My understanding was that you got prion diseases through eating
} } contaminated samples, hopefully uncommon in the world's em prep rooms. Is
} } there an airborne transmission route?
}
} The most common form of transmission for toxoplasma gondii is from eating cat
} feces but I am sure no-one would admit to eating that!
}
}
}
Personally I find a Snickers more satisfying. I have
been told oth by my Vet and our OB/GYN that it is possible to contract the
organism through the dust kicked up while kitty uses his litterbox not to
mention when we have to change it.
{} {} {} {} {} {} {} {} {} } {} {} {} {} {} {} {} {} {} {} {}
Scott Whittaker Ph 352-392-1295
ICBR EM Core Lab Fax 352-846-0251
University of Florida e-mail sdw-at-biotech.ufl.edu
www http://www.biotech.ufl.edu/~emcl/





From: scott_l-at-unin1.unorth.ac.za
Date: Wed, 20 Mar 1996 07:39:54 -0200
Subject: TEM and SEM training courses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1384845499.69898-at-msmail.tmc.tulane.edu}

Hello from the southern tip of Africa

I am emploted at an EM facility at the University of the North, South Africa.
We have been running an EM course for postgraduate students for the past 3
years. It is part of a techniques course for fourth year students in the
biological sciences, and serves as an introductory course. Following their
training, many students include ultrastructural studies as part of their
projects

I am quite eager to year of institutions that are involved in EM training.
In particular, I am interested in course content, training facilities (in-
cluding instruments), and whether or not such a course improved past
students chances of employment or to get sponsorship for futher studies.

Leon Scott
EM Unit
Univ. of the North
Private Bag X1106
Sovenga
Republic of South Africa
tel 27 152 2682957
fax 27 152 2683156/2682893




From: CAA3045-at-ritvax.isc.rit.edu (Ciprian A. Almonte)
Date: Wed, 20 Mar 1996 00:21:20 -0500
Subject: Student looking for summer job

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Fellow Microscopists,
I'm currently a biomedical photographic student at the Rochester
Institute of Technology. I have a BS in Biology and I'm currently
pursuing a second degree in Biomedical Photographic with a concentration in
photomicrography. My career objective is to apply my knowledge in biology
and biophotography to the documentation of medicine and the biological
sciences. You can view my resume and some of my images in my web site at
http://www.isc.rit.edu/~caa3045.
The biomedical photographic Communication department at RIT has
exposed me to many photography techniques. I am experience with many
formats of films processes, black-and-white, and color printing,
photomacrography, photomicrography, darkfield, brighfield, nomarski,
phase-contrast, polarizing, Rheinberg, SEM, and fluorescence microscopy.
Also, I have a great interest and hands on experience in digital
photography and computer (Macintosh major area of strength).
Also, I have a lot of experience with several immnunocytochesmistry
techniques, and the Patch-clamp techniques.
If you think I may contribute to your department send me an e-mail
to "caa3045-at-rit.edu" or "almonte-at-medcolpa.edu"
Thanks,

__________________________________________________________
Ciprian A. Almonte
Rochester Institue of Technology
Biomedical Photographic Communications
Rochester, NY 14623-5603
Visit my web site at http://www.isc.rit.edu/~caa3045/
__________________________________________________________






From: Gejing Li :      gjinli-at-umich.edu
Date: Wed, 20 Mar 1996 10:31:51 -0500 (EST)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {199603201324.IAA08945-at-IndyNet.indy.net}
To: Colin MacRae {C.Macrae-at-minerals.csiro.au} ,
Microscopy list {microscopy-at-Sparc5.Microscopy.Com}

Please add my name on your mailing list. Thanks.




From: kaurin-at-rmslab.rockefeller.edu
Date: Wed, 20 Mar 1996 11:05:40 EST
Subject: Web employment listings?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Could someone direct me to the website that has jobs/resume postings?
Thanks, Shelley Landon Kaurin





From: RobertCO2-at-aol.com
Date: Wed, 20 Mar 1996 12:18:21 -0500
Subject: subsribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am interested in subscribing to the EM news group.
Robert Sherman
"robertco2-at-aol.com"




From: howard-at-cshl.org (Tamara Howard in Cold Spring Harbor Laboratory)
Date: Wed, 20 Mar 1996 11:17:38 -0500
Subject: TEM:Eppendorfs & LR White?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9603201611.AA29625-at-phage.cshl.org}
X-Sender: howard-at-phage.cshl.org (Unverified)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Quick question: are Eppendorfs air-tight enough to use as embedding
capsules for LR White? I have cell pellets to embed and would rather not
disrupt them if I can avoid doing it.
Thanks!
Tamara






From: alan-at-macro.mse.uiuc.edu (Alan E. Lucero)
Date: Wed, 20 Mar 1996 13:55:47 -0600
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



unsubscribe




From: RNBALDUC-at-ARCRIDE.EDU.AR
Date: Wed, 20 Mar 1996 17:39 -0300
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Microscopists

The Laboratory of Electron Microscopy, School of Bioengineering,U.N.E.R
Argentina, is look for SEM to received it in donation or a very low price.
Only is required the technical information of your " work state", the model and
year of construction.
We pay all the costs of shipment
Please, for any questions contact me in

RNBALDUC-at-arcride.edu.ar

thanks in advance.





From: Katherine.S.Connolly-at-Dartmouth.EDU (Katherine S. Connolly)
Date: 20 Mar 96 16:25:16 EST
Subject: TEM:Eppendorfs+LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential

No plastic capsule or tube that I know of is air-tight enough for LR White.You
may double emded with gelatin and get away with a plastic capsule.
Kate Connolly




From: masumura-at-anvil.nrl.navy.mil (bob masumura)
Date: Wed, 20 Mar 1996 09:43:23 -0500
Subject: help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {v02130505ad75c94a74bf-at-[132.250.184.4]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"








From: Richard E. Edelmann :      edelmare-at-CASMAIL.MUOHIO.EDU
Date: Wed, 20 Mar 1996 14:02:27 -500
Subject: TMV Particles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


O.k. does anyone out there know where to get a vial of TMV
(Tobbacco Mossaic Virus) from for use as a mag standard and
particulate test specimen?

Carolina Biological is the only source I have come across but they
sell it as a whole kit for learning about Plant viruses, and I just
want the virus particles.





From: RMacKay :      RMACKAY-at-AC.DAL.CA
Date: Wed, 20 Mar 1996 11:22:08 +0000
Subject: EMPA garnet analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603201527.LAA19788-at-Snoopy.UCIS.Dal.Ca}
Comments: Authenticated sender is {RMACKAY-at-ac.dal.ca}


A colleague and I were discussing microprobe analysis of garnets and
he remarked that their analysis of Almandine often produce high totals,
a phenomenon I have also observed with high Fe garnets. I know that
others have observed this as well but has anyone come up with an
explanation ?

Bob MacKay
Robert MacKay
Department of Earth Sciences
Dalhousie University
Halifax, Nova Scotia, Canada
B3H 3J5
Tel: 902 494-7087
e-mail rmackay-at-ac.dal.ca




From: W.L. Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Wed, 20 Mar 1996 16:40:56 EST
Subject: Creutzfeld Jakob

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For those of us following this thread, the following statement issued
today by USDA-APHIS will be of interest.

} The British Ministry of Health and Agriculture have announced today that
} they may have associated CJD to BSE - an announcement was made to the
} House of Commons today.

I suppose in the future, we may be refusing to allow bovine tissues into
our labs.

-=W.L. Steffens=-
Department of Veterinary Pathology
College of Veterinary Medicine
University of Georgia




From: Chandra S. Pande :      Pande-at-anvil.nrl.navy.mil
Date: Wed, Mar 20, 1996 9:55 AM
Subject: POST DOCTORAL POSITONS AT NAVAL RESEARCH LABORATORY,WASH DC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603201429.IAA09354-at-Sparc5.Microscopy.Com}

Post doctoral positions are available at the Naval Research
Laboratory in the area of materials characterization of
superconductors for U.S. citizens with a Ph.D. in metallurgy,
physics or chemistry. Techniques utilized are TEM, SEM, SAM and
optical microscopy. Most modern instruments in these fields such
as a new CM-30 TEM are available including access to large scale
computational facilities Previous experience in superconductivity
is not required but expertise in transmission electron microscopy
is necessary. For further details, please contact:
Dr. Chandra Pande
Code 6320 tel:(202).767-2744
Naval Research Laboratory fax:(202) 767-2623
Washington, DC 20375-5343 e-mail:pande-at-anvil.nrl.navy.mil



Chandra S.Pande email: pande-at-anvil.nrl.navy.mil
phone: (202) 767-2744
fax: (202) 767-2623
Mailing address:
Code 6325
Naval Research Lab
Washington DC 20375 USA





From: Jay Jerome :      jjerome-at-bgsm.edu
Date: Wed, 20 Mar 1996 18:10:12 -0500 (EST)
Subject: Re: TEM:Eppendorfs & LR White?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






On Wed, 20 Mar 1996, Tamara Howard in Cold Spring Harbor Laboratory wrote:

} Quick question: are Eppendorfs air-tight enough to use as embedding
} capsules for LR White? I have cell pellets to embed and would rather not
} disrupt them if I can avoid doing it.
} Thanks!
} Tamara
}

We routinely embed material in eppendorfs. They work fine.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-bgsm.edu *
**************************************************************

}
}




From: krogers-at-ecn.purdue.edu (Kirk Rogers)
Date: Wed, 20 Mar 1996 13:46:21 -0500
Subject: Image Analysis Short Course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Forwarded from the NIH-Image list-

Thought many might be interested.

} Date: Wed, 20 Mar 1996 08:37:18 -0600 (CST)
} Precedence: bulk
} From: DrJohnRuss-at-aol.com
} To: Multiple recipients of list {nih-image-at-Soils.Umn.EDU}
} Subject: Image Analysis Short Course
} X-Comment: NIH Image Distribution List
}
} Apologies to those who have already read this. I posted it several months ago
} but from the number of phone and e-mail inquiries it seems worthwhile to put
} it up again.
}
} The intensive short course on image analysis and stereology that North
} Carolina State University has presented for many years will be offered three
} times in 1996, twice in Raleigh, NC (May 16-18 and May 20-22) and once in
} Denmark (June 12-15). The textbook for the course is "The Image Processing
} Handbook, 2nd Edition" by John C. Russ. Attendees also receive a CD-ROM
} (usable on both Macintosh and Windows) containing many images and a
} comprehensive set of plug-in modules for Adobe Photoshop that implement all
} of the various processing and measurement algorithms discussed in the book,
} with a detailed graphic step-by-step tutorial on disk. The faculty for the
} course include John Russ (NCSU), Robert Dehoff (Univ. of Florida) and
} Jeanette Norden (Vanderbilt Univ). This is a good opportunity to learn the
} fundamentals of quantitative image analysis as well as get answers to
} specific problems, understand how and when to use various methods and
} algorithms, and how to interpret results in a variety of fields including
} materials science, geology, biology and medicine, forensics, food science,
} etc. The course includes hands-on laboratories and the opportunity to work
} with attendees' images.
}
} For more details, you may request the printed brochure by calling Belinda
} Niedwick at NCSU Dept. of Continuing and Professional Education,
} 919-515-8185. There is also a web site with detailed course description and
} on-line registration at http://vims.ncsu.edu/matsci/IPCourse.html
}
} Thanks for your interest.
} John Russ
}

-Kirk
_____________________________________________
Kirk Rogers krogers-at-materials.ecn.purdue.edu
OR kirk.a.rogers.1-at-purdue.edu
Purdue University, School of Materials Science and Engineering,
1289 MSEE building, W. Lafayette, IN 47907-1289
OFFICE: 317-494-8751 FAX: 317-494-1204
http://materials.ecn.purdue.edu/~krogers






From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Wed, 20 Mar 1996 17:06:18 GMT
Subject: Re: TEM:Eppendorfs & LR White?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Quick question: are Eppendorfs air-tight enough to use as embedding
} capsules for LR White? I have cell pellets to embed and would rather not
} disrupt them if I can avoid doing it.
} Thanks!
} Tamara
}
}
} } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } }

Yes, as long as you nearly fill the tube with resin to exclude as much air
as possible
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-





From: dianavd-at-eye.usyd.edu.au (Diana van Driel)
Date: Thu, 21 Mar 1996 14:42:04 +1100
Subject: Re: TEM:Eppendorfs & LR White?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: diana-at-pc-0.eye.usyd.edu.au
Message-Id: {v01540b01ad76802eadd5-at-[129.78.203.31]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Quick question: are Eppendorfs air-tight enough to use as embedding
} capsules for LR White? I have cell pellets to embed and would rather not
} disrupt them if I can avoid doing it.
} Thanks!
} Tamara


They're fine, though as always happens there will possibly be a squishy bit
at the top of the tube where the LRW hasn't hardened.

Diana van Driel
Dept Ophthalmology
Sydney University
AUSTRALIA 2006






From: nano-at-ns1.LIHTI.ORG (Nanoprobes Inc.)
Date: Wed, 20 Mar 1996 17:21:42 -0500
Subject: Re: Web employment listings?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603202220.RAA28105-at-lihti.org}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Could someone direct me to the website that has jobs/resume postings?
} Thanks, Shelley Landon KaurinDear Shelley:

The new search engine, ALTA VISTA, is good for this type of thing. In
addition to a vast text index of web pages, it can also search
usenet-including the newsgroups where most jobs are posted. Address is
http://www.altavista.digital.com; if you have problems (some browsers
return a blank screen), we have a direct link to it from our "Links" page
(http://www.tiac.net/users/everlast/nano/Links.html, in the "Internet and
WWW directories" section. Try a simple query with something like "Jobs
offered microscopy" with the search field set to "Usenet".

On the Web, try Career Mosaic (http://www.careermosaic.com), which indexes
many of the jobs offered newsgroups. This has links to places to post
resumes as well.

Regards,

Rick Powell

*****************************************************************
* NANOPROBES, Incorporated *
* 25 East Loop Road, Suite 124, Stony Brook, NY 11790-3350, USA *
* *
* http://www.tiac.net/users/everlast/nano/home.html *
*****************************************************************






From: William Tivol :      tivol-at-wadsworth.org
Date: Wed, 20 Mar 1996 17:03:43 -0500 (EST)
Subject: Re: Angle measurement of single crystal faces

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} I am using an In_lens Field Emission SEM to image various single crystals.
} Can anyone suggest a technique for measuring the angles between single
} crystal faces? (Since the IFESEM is an in-lens microscope use of a double
} detector system is impossible)
}
} Is it possible to take secondary electron stereo pair images and measure the
} true crystal face angles?
}
Dear Colin,
It is certainly possible to reconstruct the 3D volume from
a stereo pair, but that is not very accurate and you would need the appro-
priate software to perform the calculation. If you have a double-tilt or
tilt-rotation specimen holder and can manipulate the specimen until one of
the faces is normal to the beam, then re-orient so that the second face is
normal to the beam, then you can get the angle between the faces more accu-
rately if you can record angular positions for the two orientations.
Yours,
Bill Tivol




From: george.braybrook-at-ualberta.ca (George Braybrook)
Date: Wed, 20 Mar 1996 16:37:45 -0700
Subject: Re: Oil Sample

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mark,

If you can sputter coat the sample with gold, you can probably get
them to pump down well enough to look at them in the SEM. We used to look
at tar sands samples with a Cambridge S250 turbo SEM after they were placed
in a vacuum dessicator for a few days to remove the light fraction of the
oil. However you will inevitably get some contamination of the column
when the beam heats up the oil, especially during x-ray analysis, so plan
to do a column clean shortly after running the oil samples.
The best way to run these types of samples is with a cryo stage.
Freeze the little buggers so they can't mess up your aperatures!!!



} I have an Amray 1600 Turbo SEM, I have a question about a sample
}
} that I would like to observe. The sample is an oil that a plant
}
} uses for a dosing machine. There are visible spots in the oil that we
}
} suspect may be Hg or Na, or the oil may be a synthetic oil, one that was
}
} not supposed to be used in the first place. I was asked if it is possible
}
} to examine a smear of the oil under SEM/EDX and identify the spots in it,
}
} but I'm thinking no because of the wet oil contaminating my sample chamber,
}
} and also the temperature of the electron beam may volatile the mercury.
}
} Is it not possible to observe this oil with my instrument? Are there any
}
} suggestions that I may follow?
}
}
} Mark Darus
} General Electric Co.
}
} Darus-at-cle.dnet.ge.com

Cheers
:)
George
Department of Earth and Atmospheric Sciences
University of Alberta
Edmonton, Alberta T6G 2E3
Canada
ph: 403-492-5746
fax: 403-492-2030






From: Pecz Bela :      pecz-at-falcon.mufi.hu
Date: Thu, 21 Mar 1996 08:46:23 +0100
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please put my name onto your mailing list (microscopy news).
Thanks in advance, Bela Pecz
21st March 1996


-----------------------------------------
Dr. Bela Pecz
Research Institute for Technical Physics
H-1325 Budapest, POBox 76
Hungary
phone: 36-1-1698-961
fax: 36-1-1698-037
E-Mail: pecz-at-falcon.mufi.hu
-----------------------------------------





From: Stephan Helfer :      S.Helfer-at-rbge.org.uk
Date: Thu, 21 Mar 1996 09:42:01 BST
Subject: Re: TEM:Eppendorfs & LR White?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

microscopy-at-Sparc5.Microscopy.Com

Dear Tamara
I have used LR White in three sizes of Eppendorfs (0.5, 1 & 2ml or
thereabouts). Air did not seem to be a problem, but the plastic used
in the tubes must have reacted with the resin. The smallest size tube
resulted in strangely shaped blocks, the bigger ones seemed ok. I
didn't persue the matter, but I could imagine that coating the inside
of the tubes with gelatine might help preventing any reaction (I also
used gelatine capsules inside Eppendorfs which worked very well).
Hope this helps

Yours sincerely

Dr Stephan Helfer, SSO
Mycologist / Plant Pathologist

Royal Botanic Garden Edinburgh, Inverleith Row, EDINBURGH EH3 5LR,
Scotland UK

http://www.rbge.org.uk

phone: +44 (0)131 552 7171 ext 280
or +44 (0)131 459 0446-280 (direct digital VoiceMail line)
fax: +44 (0)131 552 0382
============================
1896
* BRITISH
* MYCOLOGICAL
* SOCIETY
1996
A century of fungal science
============================




From: wagner-at-natlab.research.philips.com
Date: Thu, 21 Mar 1996 13:19:16 +0100 (MET)
Subject: subscribe?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I'd like to subscribe to the microscopy dicussion list,

Thanks,

Raymond Wagner.

-----------------------------------------------------------------
Raymond Wagner wagner-at-natlab.research.philips.com
Philips Research Laboratories Building WY 4.53
Prof. Holstlaan 4 Mailbox WY 42
5656 AA Eindhoven Phone +31 40 27 44573
The Netherlands Fax +31 40 27 43478




From: dago-at-odyssee.net (Dagoberto Rodriguez)
Date: Thu, 21 Mar 1996 08:56:29 -0500
Subject: Re: Oil Sample

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would filter the sample trough a membrane filter (Millipore?) and wash
with a suitable solvent (n=hexane?) to eliminate the oil. Then dry and
carbom coat.
The residue in the membrane should be easily observed in SEM.
Dagoberto Rodriguez
-------------------------

} I have an Amray 1600 Turbo SEM, I have a question about a sample
}
} that I would like to observe. The sample is an oil that a plant
}
} uses for a dosing machine. There are visible spots in the oil that we
}
} suspect may be Hg or Na, or the oil may be a synthetic oil, one that was
}
} not supposed to be used in the first place. I was asked if it is possible
}
} to examine a smear of the oil under SEM/EDX and identify the spots in it,
}
} but I'm thinking no because of the wet oil contaminating my sample chamber,
}
} and also the temperature of the electron beam may volatile the mercury.
}
} Is it not possible to observe this oil with my instrument? Are there any
}
} suggestions that I may follow?
}
}
} Mark Darus
} General Electric Co.
}
} Darus-at-cle.dnet.ge.com
}
}





From: howard-at-cshl.org (Tamara Howard in Cold Spring Harbor Laboratory)
Date: Thu, 21 Mar 1996 09:26:19 -0500
Subject: TEM:LRW/Eppendorfs revisited

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9603211420.AA17266-at-phage.cshl.org}
X-Sender: howard-at-phage.cshl.org
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Thanks to everyone who responded to my call for help. In case anyone is
interested, the replies split about 70/30 in favor of Eppendorfs, but
several people said they'd never been able to get them to work for LRW.
There were a few who suggested using gelatin capsules with the Eppendorfs -
as additional air barriers. I'll see how the plain tubes work, then try the
elaborate set-ups as needed.
Thanks again!
Tamara






From: HENRY P ADAMS :      hadams-at-nmsu.edu
Date: Thu, 21 Mar 1996 08:26:14 -0700 (MST)
Subject: Re: TEM:Eppendorfs & LR White?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Wed, 20 Mar 1996, Tamara Howard in Cold Spring Harbor Laboratory wrote:

} Quick question: are Eppendorfs air-tight enough to use as embedding
} capsules for LR White? I have cell pellets to embed and would rather not
} disrupt them if I can avoid doing it.
} Thanks!
} Tamara
Tamara: it works ok, but removing them from the eppenddorfs can be tricky
sometimes. Also, if your pellets are too dense, poor infiltration and
polymerization can be a problem.
Why not embed them in agar first in the tubes then transfer them..} } }
Hank Adams, NM






From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Thu, 21 Mar 1996 12:28:30 -0500 (EST)
Subject: actin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am startint to look at tapeworms in the Trypanorhynch group and was
wondering if someone has a suggestion for a good fix for the preservation
of "actin". I normally use 2.5% glut in cacodylate buffer. Will this be
alright to use? I greatly appreciate any help.

Peace,

Phil Rutledge




From: js_vetrano-at-ccmail.pnl.gov
Date: Thu, 21 Mar 1996 07:55 -0800 (PST)
Subject: List server

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear madam/sir

I read that you run a list server for microscopy. I believe Larry Thomas from
our lab subscribed but now that he has retired I would like to begin receiving
this service. Do I need to do anything?

Regards,

John Vetrano
SUBSCRIBE
js_vetrano-at-pnl.gov




From: HENRY P ADAMS :      hadams-at-nmsu.edu
Date: Thu, 21 Mar 1996 12:09:24 -0700 (MST)
Subject: Variable pressure SEM: fresh plant tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We recently acquired a variable pressure sem (Hitachi S3200N) with a Fullam
cold-stage, a researcher wants to image fresh leaves and is particularily
interested in the status (open vs closed) of the stomates. Does anyone
have any suggestions as to temp, pressure, Kv, beam current, etc?

TIA,
Hank Adam






From: baskin-at-biosci.mbp.missouri.edu (Tobias Baskin)
Date: Thu, 21 Mar 1996 14:47:27 -0600
Subject: Re: TEM:LRW/Eppendorfs revisited

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d07ad777008175c-at-[128.206.15.200]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Greetings,
Tamara wrote:

} Thanks to everyone who responded to my call for help. In case anyone is
} interested, the replies split about 70/30 in favor of Eppendorfs, but
} several people said they'd never been able to get them to work for LRW.
[snip]
Just wanted to point out that the term "Eppendorf tube" is a bit
like "whiskey", understood instantly, but in fact refering to a wide range
of product. So, "ep tubes" are made by gazillions of companies and
presumably out of, well, dozens, of different kinds of plastics. Might be
worth finding out from someone in the successful camp exactly where they
bought their ep tubes.

Tobias

- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
___ ____ ^ ____ _____ Tobias I. Baskin
/ \ / / \ / \ / University of Missouri
/ | / / \ / / Biological Sciences
/___ / /__ /_____\ / /__ 109 Tucker Hall
/ / / \ ( / Columbia, MO 65211 USA
/ / / \ \ / voice: 573-882-0173
/ /____ / \ \____/ /_____ fax: 573-882-0123






From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Thu, 21 Mar 1996 09:00:32 GMT
Subject: Neg. Processor Summary (long)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Here is a summary of the processor discussion. Contact the person who posed
the question to get the responses that didn't get posted to the list. In the
future, we maintain a web site in which we archive most of the discussions
for future reference. Go to the web site listed at the bottom of this page
and look in "Tips & Tricks". We have added a new search feature and would
welcome any input.




At 01:29 PM 3/20/96 -0500, you wrote:
} Not long ago, photo processors were discussed by this group. Of course
} as soon as the last message on this subject went into my trash, our
} negative processor broke and is beyond repair. It is a vintage processor
} by Hope (Model 152 B&W13-10), and spare parts are not anymore available.
} Also it was not always obvious in the discussion, if the subject was
} negative or positive processors. Please, if anyone has a summary of the
} results, I would appreciate to receive a copy by e-mail. Of course, I
} welcome any new information as well.
}
} Thanks very much,
}
} Hasso Weiland
} Alcoa Technical Cneter
} Alcoa Center, PA 15069
}
---------------------------------------------
Considering that photography is still so important to our profession, it is
likely that in this group are a few critical individuals with personal
experiencee using dry-to-dry black and white photographic processors. Would
anyone care to recommend a counter-top processor that they are happy with? I
wouldn't object to knowing which ones to stay away from either.

Many thanks,

Doug Keene
Shriners Hospital for Crippled Children
Portland, Oregon

--------------------------------------------------
} We have
been using the Ilford 2150 RC for over a year, maybe two by now.
The results are quite good, however there have been mechanical problems with
it from the get go. We had to replace bearings early on and have had to
dismantle and re-assemble the rollers over and over to keep it from
squeaking and sqealing and grinding. So it has been a mixed bag for us. We
do about 10,000 prints a year and it has been a real time-saver too
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-

------------------------------------------------------

The best processor out there is the Mohrpro. We have been using one
for about 8 years and recommend it when ever we can. It is easy to
use and maintain, 2 min. dry to dry, permanent (not stabilized) and
uses RC paper and Kodak Fixer. We do purchase the developer from
Mohr. The current cost for paper up to 8" width is $4295, the
Mohrpro8. and 14" width $4800, the Mohrpro14. (They do sell
rebuilt machines as well). They stand behind their product and I can
only say great things about Mohr.
For information contact:
Bob or Jim Jackson (tell them Linda at Loyola sent you!)
Mohr Enterprises
65 E. Palatine Rd. Suite 103
Prospect Heights, Il 60070
1-847-465-0048

Linda Fox Loyola University Medical School
lfox1-at-wpo.it.luc.edu

----------------------------------------------------------------

Dear Doug:
We've had our Mohr Pro8 for about 9 months and love it. This model came
highly recommended by other users in the Boston area who had used them
much longer than us. We now wonder how we put up with the tray method
for so long! If you want more info, E-mail or call me.
Don Gantz
Boston Univ School of Medicine
gantz-at-med-biophd.bu.edu
617-638-4017

PS. AGFA also makes one but have no details.

----------------------------------------------------------------

I have used an Agfa system, chemistry and paper for 12 years. The processor
was not sufficiently robust nor reliable for us. In any case it is no longer
available. The Agfa Variable Contrast Premium paper is my favorite emulsion
for scientific work. I recommend that you consider the Durst Printo system.
It is modular so you can build it to suit your own needs. It is my choice of
currently available systems.

: "Larry D. Ackerman" {mishot-at-itsa.ucsf.edu}
-------------------------------------------------------------------

At 10:12 AM 3/7/96 -0600, you wrote:
} The best processor out there is the Mohrpro. We have been using one
} for about 8 years and recommend it when ever we can. It is easy to
} use and maintain, 2 min. dry to dry, permanent (not stabilized) and
} uses RC paper and Kodak Fixer.

I must respectfully disagree. I was personally responsible for implementing
both an Ilford 2150 and a Mohr Pro 8" Model in a multiuser lab. While I
liked the Mohr and our users were mostly happy with it, from an ease of use
and maint. point of view, I would have to go with the Ilford.

The 2150 ran flawlessly for several years, then needed minor (in-house)
maintenance - float sensors went bad. It will process up to 20" wide sheet,
in 90 secs. dry-dry, and the rollers stay submerged in the chemistry, which
seems to keep them cleaner (no dried on chemicals to deal with).

The Mohr has some rollers partially exposed, which often was a source of
roller marks, and were more difficult to clean. We used it primarily for
negatives (TEM & SEM). It performed well, just not as smoothly as the Ilford.

Overall, I had very good support from both Ilford and Mohr (though neither
produced problems that really tested the waters). I would recommend both,
but for different apps - Ilford for RC Prints (does not do film), and the
Mohr for negs. With the caveat that the Mohr will require more hands-on time.

JCL
=========================================================
James C. Long
Manager/Materials Analysis Lab
Electrosource, Inc.
512-445-6606
jlong-at-electrosource.com

----------------------------------------------------------






} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {

Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 904-392-1295
ICBR EM Core Lab fax 904-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/





From: gwisler-at-asrr.arsusda.gov
Date: Thu, 21 Mar 1996 16:56:47 -0500 (EST)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sir: We have a dissassembled Siemens ELMISKOP 102 TEM which is
available free of charge for the taking. It was purchased in 1972, and
was working just fine when we moved it. We have a newer scope that we
like better. Please respond if you are interested.
G. C. Wisler
USDA-ARS
Salinas, Ca 93905





From: gwisler-at-asrr.arsusda.gov
Date: Thu, 21 Mar 1996 17:21:06 -0500 (EST)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Used TEM available:
1973 Siemens ELMISKOP 102. Was in fine working condition when it was
retired. Has been dissassembled and is in storage.
Interested persons contact:
Tom Nelson
USDA-ARS
1636 E. Alisal St.
Salinas, CA 93905




From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 21 Mar 1996 17:12:36 -0500 (EST)
Subject: Re: TEM safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Just a short question. What about health risks for TEM users (and
} sample preparation) when expecting a baby?
}
Dear Francisca,
The MSA Technologists Forum put together a short book published
by San Francisco Press titled Electron Microscopy Safety Handbook, VC
Barber & J A Mascorro editors. I didn't see any referrence to pregnant
workers in the Table of Contents, but the book is very useful and inex-
pensive. I ordered my copy in 1994 from

San Francisco Press, Inc.
Box 426800
San Francisco CA 94142-6800
USA
as
Electron Microscopy Safety Handbook, 2nd edition
ISBN # 0-911302-72-7

For radiation, the standard for pregnant workers is an order of
magnitude lower than for other workers, and I'd apply the same reasoning
for chemical hazards, except that I'd be even more cautious for critical
periods of pregnancy when specific organ systems are developing. Congra-
tulations and good luck.
Yours,
Bill Tivol




From: zhiyu wang :      zhiyu-at-hawaii.edu
Date: Thu, 21 Mar 1996 08:44:54 -1000
Subject: Re: sectioning teeth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English

Hi, Jean:

I worked in a project dealt with ultrasection of fish otolith for several
years. My purpose was to get thin section in {50nm for EELS analysis.
Fish otolith is a piece of rock in fish's inner ear and very hard and
fragle. I have been succesfull in getting such ultrasection in Ultracut
E machine. Few technological points may help you to get better
sections.

Resin's hardeness can not macth the hardeness of otolith no matter what
recipes you use. Also the role of resin is for hold the sample instead
of embedding. When knife touch sample, the sample get a moment tending
to rotate, thus a gap between sample and resin will occur. Also sample
can not be cut due to the elastisity of resin but none of sample. After
few cutting cycles, the sample may be cut in thick section. My method
for these problems is as following:

After a few cutting, take block out of machine, put a drop of Spurr resin
on cutting surface (sample originally embedded in Epon resin)to fill out
the gap, put it in oven for polymerilization. Repeat above for several
times. This will have the hard sample to press its surounding of resin
and increse density of resine to have better holding result. Shortly, I
grind the block with #1200 sand paper instead of cutting. By this way, I
can get silve-gold color section in 0.2-0.5 mm squar of sample area.

Picking the sections is a difficult job. I use carbon film with copper
grid to pick the sections up. Sock the carbon film with 10%-30% alcohol
then touch the sections. Alcohol reduces water surface tension and
protect sections from damage. But only 10-30% chances can get good
sections on grid.


Hope this helps


Zhiyu Wang
Department of Biosystem Engineering
University of Hawaii at Manoa
Honolulu Hawaii
zhiyu-at-hawaii.edu




jean ross wrote:
}
} I have been working on a project with a dental grad student involving
} ultramicrotomy of undecalcified human teeth. The samples were prepared by
} slicing them into approx. 100um slices and then embedding them in Epon. They
} section just great but something happens when they're picked up on grids which
} causes just the dentin portion of the section to wrinkle. I use formvar coated
} 200 mesh copper grids, stabilized with carbon and glow discharged. I spread the
} sections out in the boat with chloroform (I have used xylene too) and have
} picked them up every differnt way I could think of. Is it my support film?
} Would a thicker section be desirable (sections are usually 90-95 nm)? It is
} only the dentin part of the section that has this problem, the Epon and
} composite resin reconstruction are perfectly flat. Does anybody have any
} suggestions?
}
} Thanks in advance,
}
} Jean Ross
} Central Microscopy Research Facility
} University of Iowa
} 85 EMRB
} Iowa City, IA 52242
} (319) 335-8142
} Web site: http://www.uiowa.edu/~cemrf




From: zhiyu wang :      zhiyu-at-hawaii.edu
Date: Thu, 21 Mar 1996 09:19:45 -1000
Subject: Re: sectioning teeth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


jean ross {jeanross-at-emiris.iaf.uiowa.edu}
Hi, Jean:

I worked in a project dealt with ultrasection of fish otolith for several
years. My purpose was to get thin section in {50nm for EELS analysis.
Fish otolith is a piece of rock in fish's inner ear and very hard and
fragle. I have been succesfull in getting such ultrasection in Ultracut
E machine. Few technological points may help you to get better
sections.

Resin's hardeness can not macth the hardeness of otolith no matter what
recipes you use. Also the role of resin is for hold the sample instead
of embedding. When knife touch sample, the sample get a moment tending
to rotate, thus a gap between sample and resin will occur. Also sample
can not be cut due to the elastisity of resin but none of sample. After
few cutting cycles, the sample may be cut in thick section. My method
for these problems is as following:

After a few cutting, take block out of machine, put a drop of Spurr resin
on cutting surface (sample originally embedded in Epon resin)to fill out
the gap, put it in oven for polymerilization. Repeat above for several
times. This will have the hard sample to press its surounding of resin
and increse density of resine to have better holding result. Shortly, I
grind the block with #1200 sand paper instead of cutting. By this way, I
can get silve-gold color section in 0.2-0.5 mm squar of sample area.

Picking the sections is a difficult job. I use carbon film with copper
grid to pick the sections up. Sock the carbon film with 10%-30% alcohol
then touch the sections. Alcohol reduces water surface tension and
protect sections from damage. But only 10-30% chances can get good
sections on grid.


Hope this helps


Zhiyu Wang
Department of Biosystem Engineering
University of Hawaii at Manoa
Honolulu Hawaii
zhiyu-at-hawaii.edu

jean ross wrote:
}
} I have been working on a project with a dental grad student involving
} ultramicrotomy of undecalcified human teeth. The samples were prepared by
} slicing them into approx. 100um slices and then embedding them in Epon. They
} section just great but something happens when they're picked up on grids which
} causes just the dentin portion of the section to wrinkle. I use formvar coated
} 200 mesh copper grids, stabilized with carbon and glow discharged. I spread the
} sections out in the boat with chloroform (I have used xylene too) and have
} picked them up every differnt way I could think of. Is it my support film?
} Would a thicker section be desirable (sections are usually 90-95 nm)? It is
} only the dentin part of the section that has this problem, the Epon and
} composite resin reconstruction are perfectly flat. Does anybody have any
} suggestions?
}
} Thanks in advance,
}
} Jean Ross
} Central Microscopy Research Facility
} University of Iowa
} 85 EMRB
} Iowa City, IA 52242
} (319) 335-8142
} Web site: http://www.uiowa.edu/~cemrf




From: zhiyu wang :      zhiyu-at-hawaii.edu
Date: Thu, 21 Mar 1996 08:43:22 -1000
Subject: Re: sectioning teeth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {3151A34A.25E8-at-hawaii.edu}
jean ross {jeanross-at-emiris.iaf.uiowa.edu}
CC: Microscopy-at-Sparc5.Microscopy.Com

Hi, Jean:

I worked in a project dealt with ultrasection of fish otolith for several
years. My purpose was to get thin section in {50nm for EELS analysis.
Fish otolith is a piece of rock in fish's inner ear and very hard and
fragle. I have been succesfull in getting such ultrasection in Ultracut
E machine. Few technological points may help you to get better
sections.

Resin's hardeness can not macth the hardeness of otolith no matter what
recipes you use. Also the role of resin is for hold the sample instead
of embedding. When knife touch sample, the sample get a moment tending
to rotate, thus a gap between sample and resin will occur. Also sample
can not be cut due to the elastisity of resin but none of sample. After
few cutting cycles, the sample may be cut in thick section. My method
for these problems is as following:

After a few cutting, take block out of machine, put a drop of Spurr resin
on cutting surface (sample originally embedded in Epon resin)to fill out
the gap, put it in oven for polymerilization. Repeat above for several
times. This will have the hard sample to press its surounding of resin
and increse density of resine to have better holding result. Shortly, I
grind the block with #1200 sand paper instead of cutting. By this way, I
can get silve-gold color section in 0.2-0.5 mm squar of sample area.

Picking the sections is a difficult job. I use carbon film with copper
grid to pick the sections up. Sock the carbon film with 10%-30% alcohol
then touch the sections. Alcohol reduces water surface tension and
protect sections from damage. But only 10-30% chances can get good
sections on grid.


Hope this helps


Zhiyu Wang
Department of Biosystem Engineering
University of Hawaii at Manoa
Honolulu Hawaii
zhiyu-at-hawaii.edu



jean ross wrote:
}
} I have been working on a project with a dental grad student involving
} ultramicrotomy of undecalcified human teeth. The samples were prepared by
} slicing them into approx. 100um slices and then embedding them in Epon. They
} section just great but something happens when they're picked up on grids which
} causes just the dentin portion of the section to wrinkle. I use formvar coated
} 200 mesh copper grids, stabilized with carbon and glow discharged. I spread the
} sections out in the boat with chloroform (I have used xylene too) and have
} picked them up every differnt way I could think of. Is it my support film?
} Would a thicker section be desirable (sections are usually 90-95 nm)? It is
} only the dentin part of the section that has this problem, the Epon and
} composite resin reconstruction are perfectly flat. Does anybody have any
} suggestions?
}
} Thanks in advance,
}
} Jean Ross
} Central Microscopy Research Facility
} University of Iowa
} 85 EMRB
} Iowa City, IA 52242
} (319) 335-8142
} Web site: http://www.uiowa.edu/~cemrf




From: paulc-at-arms.gps.caltech.edu (Paul K. Carpenter)
Date: Thu, 21 Mar 1996 15:05:48 -0800
Subject: Re: EMPA garnet analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02110102ad778983ba3a-at-[131.215.67.110]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Bob MacKay asked:

} A colleague and I were discussing microprobe analysis of garnets and
} he remarked that their analysis of Almandine often produce high totals,
} a phenomenon I have also observed with high Fe garnets. I know that
} others have observed this as well but has anyone come up with an
} explanation ?

Yes, the microprobe analysis of garnets is problematic. Actually, it is
silicate minerals of the Fe-Mg solid solution series in general that share
this. Here is a short summary of my understanding of the problem:

1. It is primarily the absorption correction component of the ZAF
algorithms that is responsible for over-correcting for x-ray absorption
thus leading typically to high totals in the analysis (i.e. for garnets
101-102, and for olivines more like 101% in the total). Here I am assuming
that end member oxide standards have been used (SiO2 for Si, MgO for Mg,
and Fe2O3 for Fe, for example). So when one uses these standards, the
analysis of a garnet or olivine yields a high total. The culprit may be
that the mass absorption coefficients are in error for Mg (in particular),
but also Si and Fe; these erroneous values lead us to the high totals due
to overcorrection using a "faulty" mac. John Donovan at Berkeley drew my
attention to this, by the way.

The problem in general is that we cannot simply adjust the mac values to
suit our needs in a particular compositional system because we may not
observe the same problem in a different system.

I don't remember the particulars, but it is also known that some factors in
the fluorescence algorithms were originally fudged to work for metals
(stainless steel), and that this optimization does not hold for silicate
systems.

So there are problems in several components of typical ZAF correction
schemes and the parameters they use. Both ZAF and Phi-rho-z schemes use
macs for absorption correction, by the way.

2. One really needs to use a standard as close in composition to the
sample to be analysed when dealing with Fe-Mg garnets and olivines. I get
good results for olivines using fayalite for Fe and an Fo90 olivine for Mg
and Si when analyzing olivines that are in the Fo70-90 range. Of course
all Fe can safely be assumed to be Fe2+ in these systems (but Fe3+ in
olivines is not unknown).

Garnets are a different story. I still observe high totals when using only
garnet standards to analyze garnet samples. This again points to mac
problems, but why we have success with olivines but not garnets is an
unknown. Garnets are fairly dense and so one wonders about density terms
in the equations; however, these terms cancel when the k-ratio is
calculated. Really well characterized end-member garnets can be hard to
find (like a pure almandine, for example).

I will just mention that even though the garnet stoichiometry seems
straightforward, that I have observed fluorine up to several thousand ppm
in grossulars, and have been told about hydrogrossular component as well.
It is possible, then, that the typical wet chemical analyses of our
standard grossulars are incomplete. Fe is present as both Fe2+ and Fe3+,
but one must be careful about making charge-balance vs. stoichiometric
assumtions to calculate Fe2/Fe3.

3. Calculation of oxygen by stoichiometry works pretty well. You should
try analyzing garnets for oxygen sometime (using garnet standards, for
example grossular), and you will see that the totals are *really* bad as a
result, compared to oxygen by stoichiometry. This points to the problems
in analyzing oxygen in general.

4. Note that garnets can exhibit chemical zoning, but that this zoning
results in the same average atomic number across this zoning due to coupled
substitution, so that backscattered-electron imaging is not as successful
at elucidating inhomogeneity as in other systems. This means that the
homogeneity of a garnet standard (natural material) is suspect until
verified by mapping or linescans.

And that is just what came to mind while I sat here for a few minutes!

Paul Carpenter


+------------------------------------------------------------+
| Paul K. Carpenter |
| Division Analytical Facility |
| Geological and Planetary Sciences MC 170-25 |
| California Institute of Technology |
| Pasadena, CA 91125 |
| 818-395-6126 (X-ray Lab) 818-568-0935 (FAX, Departmental) |
| paulc-at-arms.gps.caltech.edu |
+------------------------------------------------------------+






From: Zhiyu Wang :      zhiyu-at-hawaii.edu
Date: Thu, 21 Mar 1996 20:24:33 -1000
Subject: Re: sectioning teeth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy-at-Sparc5.Microscopy.Com


I apology for posting message few times. I was not sure if the message
can be send via Netscape.


Zhiyu Wang








From: Bingqiang Lei :      Bingqiang.Lei-at-mb.luth.se
Date: Fri, 22 Mar 1996 09:06:36 +0800
Subject: who know?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {3151FD1C.7163-at-mb.luth.se}

I need to by a computer software or databank which can deal with the some
problems we materials scientists usually encounter such as the mass diffusion,
thermal expansion and thermal conductivity. Please let me know if anybody has
such a software or knows something about that. I would appreciate any clue or
information to find those software.

Bingqiang Lei

-----------------------------------------------------
Name: Bingqiang Lei
Department: Materials and Manufacturing Engineering
University: Lulea University of Technology
Tel: (46) 920 91233
fax: (46) 920 99309
Email: Bingqiang.Lei-at-mb.luth.se
-----------------------------------------------------




From: rgwhite-at-vaxc.cc.monash.edu.au (Rosemary White)
Date: Fri, 22 Mar 1996 18:06:10 +1200
Subject: Re: TEM:LRW/Eppendorfs revisited

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Re:

} } Thanks to everyone who responded to my call for help. In case anyone is
} } interested, the replies split about 70/30 in favor of Eppendorfs, but
} } several people said they'd never been able to get them to work for LRW.
} [snip]
} Just wanted to point out that the term "Eppendorf tube" is a bit
} like "whiskey", understood instantly, but in fact refering to a wide range
} of product. So, "ep tubes" are made by gazillions of companies and
} presumably out of, well, dozens, of different kinds of plastics. Might be
} worth finding out from someone in the successful camp exactly where they
} bought their ep tubes.
}

This is a good point. For consistently good results, we use "real"
Eppendorf tubes, i.e. made by Eppendorf. We did have problems with
slightly cheaper ones once, though I've forgotten exactly what the trouble
was.

Rosemary White
__ /
Department of Ecology _/ \__/ \
and Evolutionary Biology / \
Monash University / Australia \
Clayton, Victoria 3168 \ ____ /
phone 61-3-9905 5670 \_/ \_*_/
fax 61-3-9905 5613 __
email rgwhite-at-vaxc.cc.monash.edu.au \/






From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 21 Mar 1996 17:12:36 -0500 (EST)
Subject: Re: TEM safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

id sma018161; Thu Mar 21 17:07:28 1996
Received: (from uucp-at-localhost) by alcor.wadsworth.org (8.6.11/8.6.11) id RAA181
79; Thu, 21 Mar 1996 17:07:45 -0500
Received: from alcor.wadsworth.org (alcor.wadsworth.org [199.184.16.17]) by Spar
c5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id QAA12373 for {microscopy-at-sparc5.
microscopy.com} ; Thu, 21 Mar 1996 16:10:00 -0600
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11) id QA
A12377 for dist-Microscopy; Thu, 21 Mar 1996 16:10:03 -0600
Received: from Sparc5.Microscopy.Com (sparc5.microscopy.com [206.69.208.10]) by
ormail.intel.com (8.7.4/8.7.3) with SMTP id WAA14825; Thu, 21 Mar 1996 22:08:48
-0800 (PST)
Received: from ormail.intel.com by relay.hf.intel.com with smtp
(Smail3.1.28.1 #2) id m0u001p-000qDLC; Thu, 21 Mar 96 22:08 PST




From: Stephan Helfer :      S.Helfer-at-rbge.org.uk
Date: Fri, 22 Mar 1996 11:09:21 BST
Subject: Re: TEM baby

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Francisca
When our assistant got pregnant I insisted in her leaving all work
involving hazardous chemicals to other people (mainly me). The
operation of modern EMs should not be hazardous. Basically, the
pregnant operator should feel confident that she and her baby are not
at exposed to additional hazards.
The risks of something going wrong and the TEM getting the blame
(rightly or wrongly) are not worth taking.

Yours sincerely

Dr Stephan Helfer, SSO
Mycologist / Plant Pathologist

Royal Botanic Garden Edinburgh, Inverleith Row, EDINBURGH EH3 5LR,
Scotland UK

http://www.rbge.org.uk

phone: +44 (0)131 552 7171 ext 280
or +44 (0)131 459 0446-280 (direct digital VoiceMail line)
fax: +44 (0)131 552 0382
============================
1896
* BRITISH
* MYCOLOGICAL
* SOCIETY
1996
A century of fungal science
============================




From: WARRENJ1-at-cliffy.polaroid.com
Date: Thu, 21 Mar 1996 22:08 -0400 (EDT)
Subject: NIH Image

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Where can I download a copy of NIH Image?

Thanks

John Warren
Helios Scientific Group
Polaroid Corporation




From: Kingsley H. McCrocklin :      kingsley-at-zephyr.nrlssc.navy.mil
Date: 3/22/96
Subject: nanoplast/melamine resin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This message was sent by Chameleon
-------------------------------------





From: jean ross :      jeanross-at-emiris.iaf.uiowa.edu
Date: Fri, 22 Mar 1996 07:37:16 -0600 (CST)
Subject: Re: Variable pressure SEM: fresh plant tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have an S-2460N without a cold stage and have gotten great results at 25Pa
and 18Kv with about 60uA of emission. The cold stage will make a diference but
maybe this will give you a place to start.

Jean Ross
Central Microscopy Research Facility
Univ. of Iowa
(319) 335-8142
web site: http://www.uiowa.edu/~cemrf


On Thu, 21 Mar 1996, HENRY P ADAMS wrote:

} We recently acquired a variable pressure sem (Hitachi S3200N) with a Fullam
} cold-stage, a researcher wants to image fresh leaves and is particularily
} interested in the status (open vs closed) of the stomates. Does anyone
} have any suggestions as to temp, pressure, Kv, beam current, etc?
}
} TIA,
} Hank Adam
}






From: Stuart McKernan :      mckernan-at-cems.umn.edu
Date: Fri, 22 Mar 1996 09:44:13 -0600
Subject: Re: Variable pressure SEM: fresh plant tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

LIST-at-itasca.cems.umn.edu, Microscopy-at-Sparc5.Microscopy.Com

Responding to message {Pine.SUN.3.91.960321120104.6864B-100000-at-verdi}
from HENRY P ADAMS {hadams-at-nmsu.edu} :
}
} We recently acquired a variable pressure sem (Hitachi S3200N) with a
} Fullam
} cold-stage, a researcher wants to image fresh leaves and is particularily
}
} interested in the status (open vs closed) of the stomates. Does anyone
} have any suggestions as to temp, pressure, Kv, beam current, etc?
}
} TIA,
} Hank Adam
}
On our Electroscan environmental SEM we look at plant tissue at a few
degrees above freezing. This allows us to keep them hydrated at a pressure
of 5-6 torr water vapor in the chamber. At 0 Celcius the saturated vapor
pressure of water is about 4 torr, so at pressures lower than this the
material dehydrates very quickly. We typically use fairly low kV (5-10kV)
to see more surface detail. Under these conditions the plant material can
survive for reasonably long periods of time.

Stuart McKernan
University of Minnesota Characterization Facility






From: smiller-at-umr.edu (Scott Miller)
Date: Fri, 22 Mar 1996 12:01:14 -0600
Subject: WTB:Goniometer tools for Philips EM400 TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02130503ad789a59bfa3-at-[131.151.17.146]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I would like to ask if anyone has the goniometer tool kit for a Philips 400
series TEM that they would be willing to sell. I would prefer to hear from
someone in the US, but all offers will be considered.

F. Scott Miller
Electron Microscope Lab smiller-at-umr.edu
University of Missouri-Rolla voice: 314 341 4727
Rolla, MO 65401 USA fax: 314 341 6934








From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Fri, 22 Mar 1996 08:51:58 GMT
Subject: Re: TEM:LRW/Eppendorfs revisited

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Greetings,
} Tamara wrote:
}
} } Thanks to everyone who responded to my call for help. In case anyone is
} } interested, the replies split about 70/30 in favor of Eppendorfs, but
} } several people said they'd never been able to get them to work for LRW.
} [snip]
} Just wanted to point out that the term "Eppendorf tube" is a bit
} like "whiskey", .........


What I should have said was polypropylene Eppendorf-type tubes work fine.
Fisher Scientific and USA Scientific make the ones we use. We use doggie
toenail clippers to snip off the end of the tube and then tease or cut out
the polymerized sample and remount it for trimming and sectioning. If the
whole embedding process takes place in the tube, take care to see that there
is good exchange of fluids at the bottom of the tube or else you could have
trouble.

I have never drunk any whiskey from them however.

-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-





From: kna101-at-utdallas.edu
Date: Fri, 22 Mar 1996 07:57:55 -0600 (CST)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Francisca:
The lab I work in now rents time on someone else's scope, so I
haven't asked about extra saftey precautions they take, but I used to
work in a lab with it's own TEM. Our scope was checked yearly by
environmental health and saftey to make sure that the sheilding the scope
comes with was still doing it's job. It always was, but when we had a
pregnant woman using the lab., she wore a lead apron whenever she had to
be in the scope room. She also was carefull to wear gloves when handeling
chemicals, and use a fume hood. She ended up with a very healthy, rather
big, (8 pounds, plus) baby boy. She may have taken more precaution than
she needed to but it was better than not being careful enough. Good luck.
Karen Pawlowski

On Thu, 21 Mar 1996, Francisca Peiro wrote:

} Hi all,
} Just a short question. What about health risks for TEM users (and
} sample preparation) when expecting a baby?
}
} Thank you for your suggestions.
}
} F. Peiro
} =====================================================================
} Enginyeria i Materials Electronics Tel. (34-3) 402-11-39
} (34-3) 402-11-47
} Dep. Fisica Aplicada i Electronica FAX. (34-3) 402-11-48
} Universitat de Barcelona e-mail:paqui-at-iris1.fae.ub.es
} Avg. Diagonal 645-647
} 08028 Barcelona, Catalunya
} Spain
} =====================================================================
}
}




From: William Tivol :      tivol-at-wadsworth.org
Date: Fri, 22 Mar 1996 12:52:29 -0500 (EST)
Subject: Re[3]: TEM safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} } Just a short question. What about health risks for TEM users (and
} } sample preparation) when expecting a baby?
} }

[snip]

} Is the same true for SEM and FESEM (field emission SEM)?
} - JESS MUNOZ

Dear Jess,
The chemical hazards pertain to the preparation methods and are
independent of the type of instrument (except as the prep methods are mo-
dified for a particular instrument). The radiation hazard arises from
brehmsstrahlung x-rays generated by the interaction of the stray beam
electrons with apertures, the lens column and anything else that electrons
can strike. The amount of potential hazard is dependent on both voltage
and current, but the standards for radiation exposure are the same regard-
less of instrument type. As another contributor to this thread said, mod-
ern EMs are well designed for minimizing exposure and are pretty safe. If
there is any concern, a personnel monitoring device, such as a film badge
or thermoluminescent detector can be placed at the scope for a month, and
the radiation exposure can be monitored, and/or a survey meter can be used
to get a real-time readout of the radiation field.
Yours,
Bill Tivol




From: jdonovan-at-garnet.berkeley.edu (John J. Donovan)
Date: Fri, 22 Mar 1996 17:40:04 -0700
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: jdonovan-at-seismo.berkeley.edu
Message-Id: {v02120d08ad78f8d9ee31-at-[128.32.146.27]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
X-mailer: Eudora 2.1.2



===============================================================================
John J. Donovan (510) 642-5459 (phone)
Room 301, McCone Hall (510) 643-9980 (FAX)
Department of Geology and Geophysics jdonovan-at-seismo.berkeley.edu
University of California jdonovan-at-garnet.berkeley.edu
Berkeley, CA
94720-4767
===============================================================================






From: karenw-at-ucmp1.berkeley.edu (Karen Wetmore)
Date: Fri, 22 Mar 1996 11:46:41 -0700
Subject: Re: Variable pressure SEM: fresh plant tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02130502ad78914c5d28-at-[128.32.108.67]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
X-Mailer: Eudora Pro 2.1.3

I recently helped some plant biologists obtain images of pieces of maize
leaves with the stomata partially open. We used our ElectroScan ESEM with
ElectroScan's Peltier stage (a cooling stage). The images were taken at
5kV with the stage temperature set to about 5.5 degrees C. We increased
the chamber pressure to condense water on the leaf (to about 6 Torr as I
recall), then reduced the chamber pressure to slowly evaporate the water
enough to see the leaf and capture an image. The one image I saved for my
image gallery was taken at 5.3Torr. You can see it at our web site:

http://ucmp1.Berkeley.EDU/esem/gallery.htm

I have done similar condensation/evaporation experiments on other subjects
using different stage temperatures and chamber pressures, mostly around 5C
and 5Torr. The size of the subject and how closely it conforms to the stub
surface (we have cup-shaped as well as flat stubs for the Peltier) all make
a difference. The above settings will give you a starting point for
experimenting to determine what works best for your setup.

Karen Wetmore

} We recently acquired a variable pressure sem (Hitachi S3200N) with a Fullam
} cold-stage, a researcher wants to image fresh leaves and is particularily
} interested in the status (open vs closed) of the stomates. Does anyone
} have any suggestions as to temp, pressure, Kv, beam current, etc?
}
} TIA,
} Hank Adam

*****************************************************************
Karen L. Wetmore, Ph.D.
Museum Scientist
Museum of Paleontology
1101 VLSB #4780
(510) 642-0203
University of California
fax (510) 642-1822
Berkeley, CA 94720-4780 karenw-at-ucmp1.berkeley.edu
*****************************************************************






From: jdonovan-at-garnet.berkeley.edu (John J. Donovan)
Date: Fri, 22 Mar 1996 16:56:09 -0700
Subject: Re: EMPA garnet analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: jdonovan-at-seismo.berkeley.edu
Message-Id: {v02120d04ad78bed14bd4-at-[128.32.146.27]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
X-mailer: Eudora 2.1.2


} A colleague and I were discussing microprobe analysis of garnets and
} he remarked that their analysis of Almandine often produce high totals,
} a phenomenon I have also observed with high Fe garnets. I know that
} others have observed this as well but has anyone come up with an
} explanation ?

The problems with high totals in the Si-Mg-Fe system seem to be caused by a
bad mass absorption coefficient utilized by most programs of Mg Ka absorbed
by Fe. Most software I have seen use a value tabulated (but not measured)
by Heinrich. This use of Heinrich's tabulation is apparently part of the
the cause of high totals (when extrapolating from pure oxide or end-member
compositions) for many minerals, especially those with high Mg-Fe
concentrations, such olivine and garnets.

Please note the following values (soft x-ray) quoted from Heinrich and Henke :

Heinrich Henke (1982)


Mg Ka in Si 802 859
Mg Ka in Fe 6121 5250
Si Ka in Mg 2825 2902
Si Ka in Fe 2502 2305

As you can see there is about 20% difference in the mass absorption
coefficients
for Mg ka in Fe, although the others are reasonably close. This difference
will have a significant effect on the quantitative analysis (about 1 % or
so).

I have integrated the Henke value into my software as a default table and
used pure end-member olivines as standards for garnet analyses and have not
seen problems with high totals. There may be more going on here than just a
bad MAC or two, but since this is the largest correction we make to our
data, it's worth looking at first.

john

===============================================================================
John J. Donovan (510) 642-5459 (phone)
Room 301, McCone Hall (510) 643-9980 (FAX)
Department of Geology and Geophysics jdonovan-at-seismo.berkeley.edu
University of California jdonovan-at-garnet.berkeley.edu
Berkeley, CA
94720-4767
===============================================================================






From: SALLY STOWE :      stowe-at-rsbs-central.anu.edu.au
Date: Sat, 23 Mar 1996 17:24:33 EST10
Subject: Re: Variable pressure SEM: fresh plant tissue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} We recently acquired a variable pressure sem (Hitachi S3200N) with a Fullam
} cold-stage, a researcher wants to image fresh leaves and is particularily
} interested in the status (open vs closed) of the stomates. Does anyone
} have any suggestions as to temp, pressure, Kv, beam current, etc?
}
} TIA,
} Hank Adam
}
We use a Hitachi 2250N, also with a Fullam cold-stage for the
moment. . We set to around -15-20 degrees if possible to allow
for a temperature gradient through the leaf, and work usually at
around 0.1-0.3 torr, adjusting beam current to the minimum
convenient. However we generally fast-freeze the leaves in liquid
nitrogen before transferring to the SEM stage, to try to minimize
any change in stomatal state during the transition. An alternative
is to set the frozen leaf on a block of metal (preferably steel)
cooled in LN2 - this allows 30 minutes or more of observation
before drying artifacts set in.

Sally Stowe

----------------------------------------------------------------------
Sally Stowe |Email: stowe-at-rsbs.anu.edu.au
Facility Coordinator |Post:
ANU Electron Microscopy Unit |ANUEMU (RSBS)
Ph 61 6 249 2743 |Australian National Univ.
FAX 61 6 249 4891 |Canberra,
http://online.anu.edu.au/EMU/home.htm |AUSTRALIA 0200






From: cal-at-ssnet.com (cal Montgomery)
Date: Sat, 23 Mar 1996 09:14:12 -0500
Subject: Used TEM for sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Am posting this for a someone who doos not have email at the moment.

They have a fully functioning Hitachi H-600 TEM with scanning attachment
available at a reasonable price. Also have available a ultramicrotome,
knifemaker, plus.

If interested can either email back to me at cal-at-ssnet.com or call them
directly at M.H. Systems at 419-647-6400.

Instrument is still set up and running.

Thank You,

Cal
Cal Montgomery





From: Doug Keene :      DRK-at-SHCC.ORG
Date: Sat, 23 Mar 1996 08:53:28 -0800 (PST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe please




From: Dr. L. P. Stoter :      LPS-at-teknesis.demon.co.uk
Date: Sat, 23 Mar 1996 16:41:58 +0000
Subject: Subscribe microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: (Unverified)
Message-Id: {v01510100ad79d9b911fd-at-[158.152.199.245]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Subscribe microscopy

--------------------------------------------------------------
Dr. Larry Stoter
Technesis
17, Rocks Park Road, Uckfield, E. Sussex, TN22 2AT, United Kingdom
Larry-at-teknesis.demon.co.uk
--------------------------------------------------------------






From: Jill Craig :      jcraig-at-unbc.edu
Date: Sat, 23 Mar 1996 18:28:21 -0800 (PST)
Subject: Unions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Thanks for all the input regarding unions!

In summary:

Individuals working in the UK are generally unionized and are satisfied
with what unions do for them.

In the US, most are not unionized
although some would like to be or at least see points in favor of them. Some
have had the option although most have declined it. It appears that in the
US you can choose whether to be in the union or not on an individual basis.

In Canada, most are unionized although the attitude toward the union
ranges from abivalent through sceptical and resigned to decidedly anti.


Thanks for the info.

Jill




From: rh208-at-cus.cam.ac.uk (Ray Hicks)
Date: Mon, 25 Mar 1996 10:08:14 +0000
Subject: Re: Oil Sample

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: rh208-at-pop.cus.cam.ac.uk
Message-Id: {v01530500ad78358d1fe7-at-[131.111.80.78]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi,
Further to the following, wouldn't a low-tech solution work for identifying
the contaminant/s? It should be quite easy to tell if it's sodium (solid,
reacts with water to give sodium hydroxide) or mercury (dense metallic
liquid). By mixing the oil with water and shaking for a while some of the
putative sodium would react, and you could assay the aquatic phase for
NaOH. By mixing the oil with a thinning solvent, or just by centrifuging,
you'd be able to separate mercury. If neither of these yield positive
results you're left with the third option (oil). Speak to a local
inorganic chemist, or get a book out of the library on small-scale
inorganic qualitative analysis to check out the feasibility of the above.


Ray


} I would filter the sample trough a membrane filter (Millipore?) and wash
} with a suitable solvent (n=hexane?) to eliminate the oil. Then dry and
} carbom coat.
} The residue in the membrane should be easily observed in SEM.
} Dagoberto Rodriguez
} -------------------------
}
} } I have an Amray 1600 Turbo SEM, I have a question about a sample
} }
} } that I would like to observe. The sample is an oil that a plant
} }
} } uses for a dosing machine. There are visible spots in the oil that we
} }
} } suspect may be Hg or Na, or the oil may be a synthetic oil, one that was
} }
} } not supposed to be used in the first place. I was asked if it is possible
} }
} } to examine a smear of the oil under SEM/EDX and identify the spots in it,
} }
} } but I'm thinking no because of the wet oil contaminating my sample chamber,
} }
} } and also the temperature of the electron beam may volatile the mercury.
} }
} } Is it not possible to observe this oil with my instrument? Are there any
} }
} } suggestions that I may follow?
} }
} }
} } Mark Darus
} } General Electric Co.
} }
} } Darus-at-cle.dnet.ge.com
} }
} }

Ray Hicks
________________________________________________________________________
|University of Cambridge |Tel 01223 330149 |
|Department of Medicine |Fax 01223 336846 |
|Level 5, Addenbrookes Hospital |e-mail rh208-at-cus.cam.ac.uk |
|Hills Road Cambridge |Web Page/ facsmac.med.cam.ac.uk |
|CB2 |ftp server 131.111.80.78 |
|UK | |
|_________________________________|_____________________________________|






From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 25 Mar 1996 10:07:51 +0100
Subject: answers to: need books on biology...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Here is a digest of answers to my request of titles of books on biology
laboratory activities. Some people asked me to send a synthesis of responses.
I would like to thank very much all who have contributed to this list.
Bye george
giocar-at-risc990.bologna.enea.it
www.best.com/~funsci (amateur scientist site)

---------------------------------------------------------------



but a good microscopy techniques book is "Electron Microscopy:
Principles and Techniques for Biologists" by John J. Bozzola and
Lonnie D. Russell, editors. Jones and Bartlett Publishers, Boston.
1992.












--------------------------------------------------------------------
- the followings are answers to the request about "LM: lab manuals, textbooks"

Best books I know of are John Kiernan's _Histological and
Histochemical Methods_ and the Biological Stain Commission's _Staining
Procedures_ (9th or 10th [or...?] edition). I don't have Kiernan's book to
hand, but if he doesn't respond, I can send you the correct info. (Same for
_Staining Procedures_.) I heard there was a good EM book with some
procedures written by some guy in southern Illinois...


--------------------------------------------------------------
- I know this book:

Dennis E. Ohman, Experiments in gene manipulation, Prentice Hall Inc. 1988





From: Gary Login :      glogin-at-bih.harvard.edu
Date: Mon, 25 Mar 1996 10:59:07 -0500
Subject: Re: Microwave in EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In message {Pine.LNX.3.91.960325134158.3078A-100000-at-limax.paru.cas.cz}
Nebesarova LEM Motejl writes:
} I would ask somebody for an advice how to measure by simple way the
} temperature in a specimen in a domestic microwave oven and if it's better
} to remove quickly the animal tissue specimen from the warm fixation
} solution after the microwave exposure or to leave it in for long time there.
} Many thanks in advance.
} Jana Nebesarova
} Laboratory of Electron Microscopy
} Institute of Parasitology
} Ceske Budejovice
} Czech Republic


Dear Jana: your questions are very important with respect to achieving
reproducible microwave fixation.
1. I recommend a device called a 'Fix-N-Temp' container for simple, rapid,
reliable, and inexpensive temperature measurement of specimens fixed in a
microwave oven for EM studies .

a. You can make a 'Fix-N-Temp' container by sticking a liquid crystal
temperature strip into the bottom half of a 35 mm diameter plastic tissue
culture dish (Beckton Dickinson, Lincoln Park, NJ, USA) along the inside
diameter of the dish. I recommend using the liquid crystal temperature strip
available from Owl Scientific Plastics, Woburn, MA, USA (it measures
temperatures between 35 C and 65 C- ideal for the temperature range for
microwave fixation). Alternatively, this device can be purchased ready made for
several dollars from some EM Supply companies.

b. Place your fixative (up to 5 ml) in the dish, place your tissue (1 mm3 up to
1 cm3) in the fixative and irradiate according to your fixation protocol. You
will see the temperature of the solution displayed on the strip as soon as you
remove the dish from the microwave oven. The temperature is accurate to within
5 C. The container and liquid crystal strip are reusable.

c. Other methods for measuring temperature require the use of electronic
temperature probes which are more expensive, have a slower response time, and
are in my experience less accurate. In addition, the temperature probes in
microwave ovens are known to distort the microwave fields and cause conductive
heating artifacts if placed near a tissue sample.

2. The tissue should be removed from the warmed fixative as soon as microwave
irradiation has stopped. Tissues left in warm fixative are damaged by
conductive heating from the solution. Specimens up to several cubic mm can be
fixed in less than 30 seconds in a microwave oven that has been properly
calibrated.

These issues are discussed in detail in the following two references:
1. Login GR, Dvorak AM. Methods of microwave fixation for microscopy. A review
of research and clinical applications: 1970-1992. Prog Histochem Cytochem
1994;27/4: 1-127.

2. Login GR, Dvorak AM. The Microwave Toolbook. A Practical Guide for
Microscopists (Beth Israel Hospital, Boston, 1994).

Please contact me if you would like additional information.




Gary R. Login, D.M.D., D.M.Sc.
Assistant Professor of Oral Pathology
Beth Israel Hospital
Department of Pathology
330 Brookline Avenue
Boston, Massachusetts, 02215

glogin-at- bih.harvard.edu
Telephone: 617-667-2034
Fax: 617-667-8676





From: petford-at-vax.ox.ac.uk
Date: Mon, 25 Mar 1996 09:22:19 +0000
Subject: TEM safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Francisca, I work in the Materials Department at the University of Oxford,
and carry out a lot of TEM work, especially using 400kV JEOL machines. I have
had two pregnancies whilts using the TEMs, as has another woman in my
department, all with absolutely no problems at all. We had the Radiation
Protection Office check our machines for radiation levels and they all came way
below any danger levels. The only difficulty comes when you get too big to
reach the console!!

Amanda Petford-Long.




From: dianavd-at-eye.usyd.edu.au (Diana van Driel)
Date: Mon, 25 Mar 1996 10:48:28 +1100
Subject: Re: NIH Image

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: diana-at-pc-0.eye.usyd.edu.au
Message-Id: {v01540b00ad7b8f10eaa8-at-[129.78.203.31]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Where can I download a copy of NIH Image?


NIH Image home page is at http://rsb.info.nih.gov/nih-image/. From here you
download the programme (current is 1.60b7) and manual.

Diana van Driel
Dept Ophthalmology
Sydney University
AUSTRALIA 2006






From: dianavd-at-eye.usyd.edu.au (Diana van Driel)
Date: Mon, 25 Mar 1996 10:12:30 +1100
Subject: Re: TEM baby

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: diana-at-pc-0.eye.usyd.edu.au
Message-Id: {v01540b02ad7b84d645e1-at-[129.78.203.31]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

When I was pregnant I did almost no lab work involving hazardous chemicals
(fixing, embedding and photography particulary). I cut sections, but was
careful of the resin (no sawing blocks in half). I kept EM time to a
minimum. There are regulations in Aust. stipulating what pregnant women can
be exposed to and the concentrations are MUCH lower than for general
workers, but basically steer clear of the lot if you can. Babies and the
adults they grow into are a lot more important than missing out on
processing a few specimens - someone else can do it if it's important. If
the powers that be don't agree, stand up for your babies rights and don't
give in.

Diana van Driel
Dept Ophthalmology
Sydney University
AUSTRALIA 2006






From: Len Adleman :      adleman-at-pollux.usc.edu
Date: Sun, 24 Mar 1996 12:36:21 -0800 (PST)
Subject: Please subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to subscribe:
adleman-at-cs.usc.edu
(I am unsure that this group is appropriate for my needs so
it is possible I will unsubscribe after a short trial)
Thanks




From: yimei-at-befvax.uchicago.edu
Date: Mon, 25 Mar 1996 11:50:05 EST
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






From: James S MArtin :      James.S.Martin-at-williams.edu
Date: Mon, 25 Mar 1996 12:50:31 -0500 (EST)
Subject: Re: EDS atlas of materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jesus:

The Particle Atlas is an invaluable resource for particle analysts.
Volumes 3 and 6 may contain SE micrographs and EDS spectra of use to
you. The Particle Atlas was originally printed as a six-volume hardbound
edition:

The Particle Atlas. ed. Walter McCrone. Ann Arbor, MI: Ann Arbor Science
Publishers, Inc.

The Particle Atlas is also available on CD-ROM from MicroDataware
(800/582-6624). See a review by Walter McCrone:

McCrone, W. "The Particle Atlas, Electronic Edition," American Laboratory
(April 1993): 39-44.


James Martin
Director of Analytical Services and Research
Williamstown Art Conservation Center

standard disclaimer of financial interest


On Mon, 25 Mar 1996, Jesus Munoz wrote:
n
}
} Hi all.
}
} I'm currently doing EDS (Energy Dispersive Spectroscopy, some call it
} EDX) analysis of contaminants caught in our production line. As you
} know, EDS gives the composition of a substance under analysis. But
} what's more important is for us to know/identify what the material is
} and where it came from (ex. Fe, Ni, Cr is stainless steel, Powder has
} Si & Mg).
}
} Is there a book, magazine, atlas, or anything that could give me such
} information? I'd appreciate any help you could give. Thanks..
}
} JESS MUNOZ
}
}






From: Dave King (607)857-1248 T37/257-3 Ext DEKING-at-VNET.IBM.COM
Date: 25 Mar 1996 12:46:35 EST
Subject: EDS atlas of materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603251744.LAA02166-at-Sparc5.Microscopy.Com}

To: Microscopy ListSer {Microscopy-at-MSA.Microscopy.Com}
cc: VANHART --ENDVM5 Dan VanHart 7-1262
*** Reply to note of 03/25/96 09:23

McCrone's Particle Atlas is great for materials source ID info.
The McCrone Institute is in Chicago. They have a home page at;

http://www.mcri.org

{
} { {
} } ===========} Dave King { { {
} } } ------------------------------------------------------ { { { {




From: Joseph P. Neilly 708-938-5024 :      NEILLY.JOSEPH-at-igate.pprd.abbott.com
Date: Mon, 25 Mar 1996 13:57:00 -0600 (CST)
Subject: Re:EDS atlas of materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta RANDD; Relayed; Mon, 25 Mar 1996 14:08:11 -0600
Mr-Received: by mta MCM$RAND; Relayed; Mon, 25 Mar 1996 14:08:13 -0600
Mr-Received: by mta RANDB; Relayed; Mon, 25 Mar 1996 14:08:21 -0600
Alternate-Recipient: prohibited
Disclose-Recipients: prohibited
Content-Return: prohibited

Jess,

You could try the McCrone Particle Atlas. It is no longer in print but
an electronic copy is available on CD-ROM from:

MicroDataWare
2894 Tribune Ave
Hayward, CA 94542
Phone: 1-800-582-6624

The atlas contains microscopic data from a variety of materials
including optical micrographs, SEM and EDS.

Good Luck,

Joe Neilly
Abbott Laboratories
North CHicago, IL






From: Nebesarova LEM Motejl :      lem-at-paru.cas.cz
Date: Mon, 25 Mar 1996 13:53:10 +0100 (GMT+0100)
Subject: Microwave in EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would ask somebody for an advice how to measure by simple way the
temperature in a specimen in a domestic microwave oven and if it's better
to remove quickly the animal tissue specimen from the warm fixation
solution after the microwave exposure or to leave it in for long time there.
Many thanks in advance.
Jana Nebesarova
Laboratory of Electron Microscopy
Institute of Parasitology
Ceske Budejovice
Czech Republic




From: Douglas F Bowling :      al428-at-dayton.wright.edu
Date: Mon, 25 Mar 1996 15:43:19 -0500
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe




From: SveEn-at-pai.liu.se (Sverker Enestrom)
Date: Mon, 25 Mar 1996 11:12:26 +0200
Subject: Re: Stains

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bob:
Why not simply try Periodic acid-Schiff? Periodic acid will cleave the glyceryl
C-C bonds adjacent to -OH groups; the aldehydes are then demonstrated with
Schiff.
Sverker

} Does anyone know of a stain that is specific to either carbonyl or hydroxyl
} groups? If possible, I would like to find a way to visibly stain gylceryl
} monostearate on the surface of polypropylene.


*********************************************************
Sverker Enestrom M.D., Ph.D.
Department of Pathology
University of Linkoping, Sweden
Phone: +46 13 22 15 20
Fax: +46 13 13 22 57
*********************************************************






From: Francisco J. Hernandez :      fjhblasq-at-biomed.icb2.usp.br
Date: Mon, 25 Mar 1996 07:33:00 -0300 (EST)
Subject: LM: diamond knife for resin sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I want to buy a 8 mm diamond knife to cut material
embedded in historesin, but I would like some
information about it.
Does anyone have any experience with this material?
Who sells it and how much does it cost?
I will be very grateful for any help.

_____________________________________________________________________________
Francisco Javier Hernandez Blazquez * Av. Prof. Lineu Prestes, 1524
Departamento de Histologia e Embriologia * 05508-900 Sao Paulo
Instituto de Ciencias Biomedicas * e-mail fjhblazq-at-spider.usp.br
Universidade de Sao Paulo * fjhblasq-at-biomed.icb2.usp.br
______________________________________________________________________________






From: Jeffrey.Shield-at-mse.utah.edu (Jeff Shield)
Date: Mon, 25 Mar 1996 14:26:18 -0700 (MST)
Subject: Focused Ion Beam Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,

A colleague is needing compositional analysis from 10 nm layers. He is
interested in doing focused ion beam analysis considering sample prep
difficulties with TEM. Since I am not familiar with FIB, is this a viable
technique for his problem? If it is, is there anyone out there willing to
analyze his samples?

Jeff Shield





From: pwkr-at-ugcs.caltech.edu (paul wilhelm karl rothemund)
Date: Mon, 25 Mar 1996 15:15:05 -0800 (PST)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to subscribe:
rothemun-at-pollux.usc.edu

Thanks, Paul Rothemund




From: HENRY P ADAMS :      hadams-at-nmsu.edu
Date: Mon, 25 Mar 1996 13:54:02 -0700 (MST)
Subject: Re: EDS atlas of materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Mon, 25 Mar 1996, Jesus Munoz wrote:

}
} Hi all.
}
} I'm currently doing EDS (Energy Dispersive Spectroscopy, some call it
} EDX) analysis of contaminants caught in our production line. As you
} know, EDS gives the composition of a substance under analysis. But
} what's more important is for us to know/identify what the material is
} and where it came from (ex. Fe, Ni, Cr is stainless steel, Powder has
} Si & Mg).
}
} Is there a book, magazine, atlas, or anything that could give me such
} information? I'd appreciate any help you could give. Thanks..
}
} JESS MUNOZ
}
There is an atlas authored by Walter C. McCrone and John G. Delly, titled
the The PARTICLE ATLAS, a 4 volume set, however, vol.IV contains edx
spectra and SEM of particles. The other volumes deal with other aspects
of particle identification such as princibles and techniques, polarizing
microscopy, etc. They are all useful. The set we have in our lab is 2nd
ed, copyright 1973 by Ann Arbor Science Publishers, Inc., Ann Arbor,
Michigan, USA. ISBN # 0-250-40008-1. I don't know if there is a new
edition out, Walter McCrone is the father of particle identification.
Good luck, Hank Adams
}




From: Scott Williams :      scott_williams-at-pch.gc.ca
Date: 3/25/96 6:15 PM
Subject: EDS atlas of materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: prohibited
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Normal



Hi all.

I'm currently doing EDS (Energy Dispersive Spectroscopy, some call it
EDX) analysis of contaminants caught in our production line. As you
know, EDS gives the composition of a substance under analysis. But
what's more important is for us to know/identify what the material is
and where it came from (ex. Fe, Ni, Cr is stainless steel, Powder has
Si & Mg).

Is there a book, magazine, atlas, or anything that could give me such
information? I'd appreciate any help you could give. Thanks..

JESS MUNOZ


The following is an attached File item from cc:Mail. It contains
information that had to be encoded to ensure successful transmission
through various mail systems. To decode the file use the UUDECODE
program.
--------------------------------- Cut Here ---------------------------------
begin 644 RFC822.TXT



From: Scott Williams :      scott_williams-at-pch.gc.ca
Date: 3/25/96 6:15 PM
Subject: EDS atlas of materials

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

end





From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 25 Mar 1996 14:21:34 -0800 (PST)
Subject: Re: LM: diamond knife for resin sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello, We have been using the Diatome Histo-Diamond knives for several
years to cut Historesin and epoxies. These knives are very durable and a
real timesaver for extensive serial sectioning and for cutting things that
savage glass knives, like a little bone or otoconia. They've survived
sectioning up to 5 microns and cut well down to 0.5. At 0.5 microns the
sections start to look a little scratchy. They won't section a lot of
bone, someone here made an unauthorized attempt to section an untrimmed
undecalcified cochlea. The knive sectioned with lots of scratch marks
afterwards, but cut well after being sent back for resharpening.

You will want to clean the knive periodically with mild detergent for
optimal performance.

Regards,

Glen MacDonald
Research Scientist
Hearing Research Laboratories of the
Virginia Merrill Bloedel Hearing Research Center
Box 35-7923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156
glenmac-at-u.washington.edu


On Mon, 25 Mar 1996, Francisco J. Hernandez wrote:

}
} I want to buy a 8 mm diamond knife to cut material
} embedded in historesin, but I would like some
} information about it.
} Does anyone have any experience with this material?
} Who sells it and how much does it cost?
} I will be very grateful for any help.
}
} _____________________________________________________________________________
} Francisco Javier Hernandez Blazquez * Av. Prof. Lineu Prestes, 1524
} Departamento de Histologia e Embriologia * 05508-900 Sao Paulo
} Instituto de Ciencias Biomedicas * e-mail fjhblazq-at-spider.usp.br
} Universidade de Sao Paulo * fjhblasq-at-biomed.icb2.usp.br
} ______________________________________________________________________________
}
}
}





From: scott_l-at-unin1.unorth.ac.za
Date: Tue, 26 Mar 1996 12:28:02 -0200
Subject: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603260425.XAA00633-at-use.usit.net}

Are there IBM compatible PC based programmes able to assist with doing
stereology from TEM and LM micrographs. What are the hardware require-
ments?

Thank you
Leon Scott
SA




From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Tue, 26 Mar 1996 08:38:11 GMT
Subject: Re: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 12:28 PM 3/26/96 -0200, you wrote:
} Are there IBM compatible PC based programmes able to assist with doing
} stereology from TEM and LM micrographs. What are the hardware require-
} ments?
}
} Thank you
} Leon Scott
} SA
}
}
You will find an archived discussion about the presentation
of stereo pairs on the web page listed at the bottom of this message. Look
in the "Tips & Tricks" section. There was some discussion about software. I
will be more than happy to e-mail the file to you as well if you are unable
to access the page.
Cheers




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {

Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 904-392-1295
ICBR EM Core Lab fax 904-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/





From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Tue, 26 Mar 1996 08:54:37 -0500 (EST)
Subject: actin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Fellow Microscopists:

I am starting to look at tapeworms in the Trypanorynch group and was
wondering if there is a suggestion for a good fix for the preservation of
"actin"? I normally use 2.5% glut in cacodylate buffer. Will this be
O.K. or is there something better? I greatly appreciate any help.

Peace,

Phil Rutledge
8-{)




From: Elinor Solit :      cambrex-at-world.std.com
Date: Tue, 26 Mar 1996 10:48:42 -0500 (EST)
Subject: Re: (no subject)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Keith,

Send us your name and address and we'll send you a copy of the beautiful
"Microscope Book", a catalog with a lot of information. Or use our 800
#, 440-0311.

Hope all your horses are winners.

Ellie Solit
The Cambrex Group


On Mon, 25 Mar 1996, Keith Lewis Allison, Jr. wrote:

} Hello!
} I have a small horse breeding operation and am just beginning to use
} artificial insemination. I am in need of a microscope for semen
} evaluation. I would appreciate any advice you may be able to provide on
} the purchase, i.e. objective size, source, options, etc. Please keep in
} mind that I am ignorant of microscopy terminology. I do not want
} anything exceptionally fancy, but must be functional....and "tough" for
} farm use. Thanks in advance for your response!
} Keith
}
}




From: Dave King (607)857-1248 T37/257-3 Ext DEKING-at-VNET.IBM.COM
Date: 26 Mar 1996 12:24:26 EST
Subject: Polaron sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603261722.LAA03719-at-Sparc5.Microscopy.Com}

To: Microscopy ListSer {Microscopy-at-MSA.Microscopy.Com}
*** Reply to note of 03/26/96 11:50

The sputter targets I've seen are thin, and when you see them
starting to come apart, it's time to change them, if you're
concerned about minor contaminants from the base material (I
think ours is Al).

Try Ernest Fullam, SPI, etc for replacement targets.

{
} { {
} } ===========} Dave King { { {
} } } ------------------------------------------------------ { { { {




From: Lesley S. Smith :      lesleys-at-pobox.upenn.edu
Date: Tue, 26 Mar 1996 14:27:00 -0500 (EST)
Subject: For Sale!!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Zeiss TEM-900 for Sale

-Turbomolecular pump, sheet film camera, extras as well.
-asking $35,000. or make an offer.
-call 904-775-4330




From: Douglas F Bowling :      al428-at-dayton.wright.edu
Date: Tue, 26 Mar 1996 14:50:52 -0500
Subject: SEM's in High Schools

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am a high school teacher of Electronics, who also has a Amray 1700 SEM as a
Science enrichment class for Dayton City Schools in Dayton, OH.

Am interested to know if others out there have an SEM in their high schools.

Hoping to compare notes and ideas.

Thanks,
Doug




From: m.dickson-at-unsw.edu.au (melvyn dickson)
Date: Wed, 27 Mar 1996 10:03:45
Subject: Re: SEM, ESEM examination of soils

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello out there,
Local microbiologists want to look at soil samples and the resident microboal
flora. My EM lab has all the usual SEM and TEM methods on tap and are doing a
library search but if anyone out there has personal experience of

A: looking for (or at) microbes in soil (ON soil) with frozen-hydrated
specimens or

B: untreated soil in an environmental SEM.

we would be very interested to learn of your experiences.
Thanks in advance,

Mel Dickson.




From: dbd1-at-uclink4.berkeley.edu
Date: Tue, 26 Mar 1996 15:43:41 -0800 (PST)
Subject: need section counter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To all,

I am looking for a section counter accessory for the Reichert Ultra Cut E.
Anybody out there have one that is not being used?


Doug Davis
Staff Research Associate
Electron Microscope Facility
University of California
Berkeley, CA 94720
(510) 642-2085
dbd1-at-uclink4.berkeley.edu







From: Linda Fox :      lfox1-at-wpo.it.luc.edu
Date: Tue, 26 Mar 1996 09:39:47 -0600
Subject: Polaron sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s157bb1f.039-at-wpo.it.luc.edu}
X-Mailer: Novell GroupWise 4.1

Can anyone tell me when it is time to purchase a new Au target for
our Polaron E5100 Series II sputter coater? Ours seems cracked and
peeling in places. Over time it is hard to say if there has been a
difference in coating. There has been a difference in the current we
use to sputter coat. Also, if anyone knows who carries a line of
Polaron parts, that information would be helpful as well.
THANKS
Linda Fox
Loyola University Medical School
lfox1-at-wpo.it.luc.edu





From: Deborah Holmberg :      dlholmberg-at-ucdavis.edu
Date: Tue, 26 Mar 1996 10:59:43 -0800 (PST)
Subject: Re: Scanning 96, student volunteers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Thu, 7 Mar 1996, Deborah Holmberg wrote:

} Hello All,
} The Scanning 96 Meeting in Monterey, CA on 9-12 April is looking for
} students to monitor the sessions, ie. run the slide projectors, adjust
} lights and hand out information. The student will have the registration
} fee waived, in appreciation of helping. For information, please contact
} me, or the Scanning 96 office at FAMS, Inc., PO Box 832, Mahwah, NJ
} 07430-0832. E-mail: fams-at-holonet.net
} Thank you,
} Debe Holmberg
} USDA-ARS
} 916.752.9021
} 916.752-4604 (fax)
}




From: Cheri Owen :      cowen-at-cmb.biosci.wayne.edu
Date: Tue, 5 Mar 1996 12:29:18 -0500 (EST)
Subject: Used TEM for sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Just wanted to post one last time before we quit.

---------- Forwarded message ----------



Recently, I sent a message to you all about a used TEM, to see if there
was any interest. I have had some, so here is the real offer.

We have a Philips 201c TEM for sale, it is approximately 20 years old.
It is functional, but lack of use has caused the vacuum to be less than
optimal. With a little TLC this scope could be an excellent workhorse.
The investigators who own the scope would be willing to sell it for
parts, but they want it removed from their facility, not just scavenged,
as they need to make room for a new scope.

Thanks in advance

Cheri Owen
Wayne State University
Detroit Neurotrauma Institute
Detroit, MI
(313)577-4648





From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Tue, 26 Mar 1996 19:12:57 -0600
Subject: Re: EMPA: anybody ever done hair?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Does anyone out there have any experience with EMPA of hair? Preparation,
} mounting, etc? Someone here is interested in looking at heavy metals that
} thru metabolism are concentrated in the hair.The levels will be low, so
} the MDL will be a question, but I don't know how the hair will stand up to
} an electron beam.
}
} John
}
}
} John Fournelle

I just recently did some dog hair. It seemed to hold up well under
a 10kV beam. This was for imaging, so was Au/Pd coated, however.
Carbon-coating will be more interesting, but should still work. Since your
person wants heavy metals, it might be worthwhile using low kVs and going
for M-lines. Try obilgue or cross-sectioned hairs--shaft exteriors don't
seem to do so well for x-ray; least I've never gotten anything from them.
Phil Oshel

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: Leah Dobbs :      leadob-at-execpc.com
Date: Tue, 26 Mar 1996 21:39:20 -0800
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {3158D488.7F8A-at-execpc.com}

unsubscribe




From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Tue, 26 Mar 1996 16:34:52 GMT
Subject: Re: Stereology (apology)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sorry guys. I must have been absent the day they taught us
to read



} At 12:28 PM 3/26/96 -0200, you wrote:
} } Are there IBM compatible PC based programmes able to assist with doing
} } stereology from TEM and LM micrographs. What are the hardware require-
} } ments?
} }
} } Thank you
} } Leon Scott
} } SA
} }
} }
} You will find an archived discussion about the presentation
of stereo pairs on the web page listed at the bottom of this message. Look
in the "Tips & Tricks" section. There was some discussion about software. I
will be more than happy to e-mail the file to you as well if you are unable
to access the page.
} Cheers
}




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {

Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 904-392-1295
ICBR EM Core Lab fax 904-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/





From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Wed, 27 Mar 1996 00:24:51 EST
Subject: Sputter coater targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

Linda Fox wrote:

Can anyone tell me when it is time to purchase a new Au target for our
} Polaron E5100 Series II sputter coater? Ours seems cracked and
peeling in
} places. Over time it is hard to say if there has been a difference
in
} coating. There has been a difference in the current we use to
sputter
} coat. Also, if anyone knows who carries a line of Polaron parts,
that
} information would be helpful as well.

It sure does sound like you need a new cathode.

The Polaron line is now being produced by VG Microtech in the UK. The
"official" distributor for VG Microtech in the USA is Energy Beam
Sciences.

Several other firms, such as SPI Supplies are also manufacturing
replacement targets for these units. If your system takes the 57 mm
solid disc, the SPI replacement cathodes differ from the original
equipment cathodes in that the SPI version is 10 mils thick (instead of
3 or 5 mils). Don't forget that SPI offers a precious metals recycling
program, granting a 10% discount when the spent cathode is returned at
the time of purchase of the new cathode.

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Wed, 27 Mar 1996 00:23:17 EST
Subject: Embedding w/ LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

The following was posted recently:

No plastic capsule or tube that I know of is air-tight enough for LR
White.You may double emded with gelatin and get away with a plastic
capsule.
Kate Connolly

Standing in our exhibit booth, and listening to how some of our more
ingenuous customers use some of our products, I have learned that if
one takes the UV transparent SPI embedding (silicone) molds, and over
fills them slightly, and then places another (identical) mold on top,
the capillary action between the two molds really does seal out oxygen
to the point that the resin can be UV cured without worry that oxygen
will some how interfere.

Of course, after polymerization, the top mold separates easily and the
blocks are removed from the "bottom" mold.

There are a number of advantages to using this approach since it is far
more easy to properly align samples in a specific direction with
respect to what will eventually become the cutting direction.

More information about the SPI Supplies silicone embedding molds can be
found in our electronic catalog on the WWW (see below).

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: Stephan Coetzee :      stephan-at-gecko.biol.wits.ac.za
Date: Wed, 27 Mar 1996 09:13:03 GMT+2
Subject: Re: Polaron sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Linda Fox wrote:

} Can anyone tell me when it is time to purchase a new Au target for
} our Polaron E5100 Series II sputter coater? Ours seems cracked and
} peeling in places. Over time it is hard to say if there has been a
} difference in coating. There has been a difference in the current we
} use to sputter coat. Also, if anyone knows who carries a line of
} Polaron parts, that information would be helpful as well.
} THANKS

Here we using the same model. Normally there is a area which the
target suffer a higher rate of sputtering material loss. That is normally
at a area at the edge along the perimeter of the target. It eventually
leeds to a groove, exposing the target ring. When that happens we
replace ours. We buy the Au or Au/Pd in sheets locally and do it our
selves. It is cheaper and saves the time spent waiting for the
target to arrive. (If it is not in stock we wait 3-8 weeks!) Note
that the sheets can be difficult to cut, especially the Au/Pd

Hope this helps.


Stephan H Coetee
Electron Microscope Unit
Private Bag 3
Wits
2050 Stephan-at-Gecko.biol.WITS.ac.za

Tell: (011) 716 2419
Fax : (011) 339 3407




From: M.F. Butler :      mfb12-at-cus.cam.ac.uk
Date: Wed, 27 Mar 1996 08:50:52 +0000 (GMT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe please




From: Dirk.Voeste-at-rz.ruhr-uni-bochum.de
Date: Wed, 27 Mar 1996 10:41:05 +0000
Subject: TEM-fixation of waterplants

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Everyone,
I am working on a project with TEM investigations of higher waterplants.
Untill now I have only a few experiences with electronmicroscopical
analyis. In my lab we tried to fix the waterplants with large
gaslacunae in the leaves and only two to three cell layers arround
it. We have always the problem that the tissue collapse by
evacuating the leaves or embedding them into Epon, Spurr etc., when
they are not evacuated. The fixation and contrast of cytoplamsic structures
are OK.
Has anyone an idea or knows a method to solve the problem of tissue
collapsing? Is a there a paper published where this phenomen or a appropriate
method has been described for submers waterplants, especially with large
gaslacunae?

Thanks for your help!

Dirk Voeste


Dr. D. Voeste
Ruhr-Univercity Bochum
Comparative Endocrinology Research Section
ND-5/31
D-44780 Bochum
0234/7004325 (phone)
0234/7094551 (fax)
dirk.voeste-at-rz.ruhr-uni-bochum.de




From: ENERGY BEAM SCIENCES, INC :      75767.640-at-CompuServe.COM
Date: 27 Mar 96 08:19:18 EST
Subject: Re: Polaron sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Jean-Pierre Slakmon {jps-at-soquelec.com}

Linda-
The Polaron range of specimen preparation equipment for electron microscopy is
manufactured by VG Microtech in the U.K. We (Energy Beam Sciences) are the
exclusive representative for the Polaron range in the United States; Soquelec
Ltd. are the exclusive representatives in Canada. We provide authorized bench
service for current and all older models of Polaron instruments, and carry a
wide range of spare parts in inventory, including targets for all models of
sputter coaters.
The current Polaron range is described at our WWW site
(http://www.ebsciences.com/).
Please contact me directly if I can be of further assistance.
Best regards,
Steven E. Slap, Vice-President
75767,640-at-compuserve.com





From: John Gabrovsek :      gabrovj-at-cesmtp.ccf.org
Date: Wed, 27 Mar 1996 09:25:15 -0500
Subject: TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s1590bd2.019-at-cesmtp.ccf.org}
X-Mailer: Novell GroupWise 4.1

Hi everybody,
A friend of mine would like to acquire an RCA EMT electron microscope
if they are still available. If they are please get in touch directly
with him. Mr. Tom Bunch e-mail: { tbunch2-at-ix.netcom.com }
Regards,
John Gabrovsek
CCF Cleveland,Ohio





From: Richard Beanland +44 1327 356363 :      richard.beanland-at-gecm.com
Date: Wed, 27 Mar 1996 16:46:22 +0000 (GMT)
Subject: Gold sputter targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Disclose-Recipients: prohibited

Seeing Linda Fox's post just reminded me of something I found out the other
day here. Apparently the cheapest source of gold (99.9% pure) is new 'limited
edition' gold coins. Don't ask me why! One guy here had a crown about an
inch across which he then rolled to produce a 2 inch target for an industial
coater. Saved the company a packet.


Richard Beanland
GMMTL Caswell,
Towcester,
Northants NN12 8EQ.

Tel. +44 1327 356363
Fax. +44 1327 356775
Email richard.beanland.gecm.com





From: Igor Polyakov :      Igor_Polyakov-at-qmgate.arc.nasa.gov
Date: 27 Mar 1996 09:15:22 -0800
Subject: freezing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1384204335.95012-at-qmgate.arc.nasa.gov}

Subject: Time:9:42 AM
OFFICE MEMO freezing Date:3/26/96

I am looking for available information about freezing a tissue(brain) for LM
histochemistry and immunocytochemistry. Also, I need information how to
estimate the optical density of a histochemically stained sections? Are there
books, those could give me such information?
Any information would be very much appreciated.
Igor_Polyakov-at-qmgate.arc.nasa.gov






From: John Phelps :      phelps-at-enh.nist.gov
Date: Wed, 27 Mar 1996 11:41:09 -0700 (MST)
Subject: Sample prep - Glue Sources?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good day all,

We have been using Gatan "G-1" epoxy to prepare cross sections for TEM
analysis. Does anyone know what this "G-1" epoxy is? More importantly, is
this an epoxy that be purchased at, say a hardware store? Or perhaps some
one knows of a substitute that works as well or possibly better that can be
purchased locally.

thanks in advance,
John

John Phelps
NIST
Materials Reliability Division
325 Broadway
Boulder, CO 80303
ph. 303-497-7570
fax. 303-497-5030




From: Hand-at-nso1.uchc.edu (Hand,Arthur)
Date: Wed, 27 Mar 1996 12:01:07 -0500
Subject: EM Facility User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to all who responded to my request for information on 3/11/96. I have
compiled a list of fees for selected services for 12 USA facilities and 4
facilities in other countries. If you would like a copy, let me know if you
prefer to receive it by FAX or by e-mail (Mac or IBM format).

Arthur R. Hand
Central EM Facility
Univ of Connecticut Health Ctr
e-mail: hand-at-nso1.uchc.edu





From: MICHAEL KNOTTS :      ph281mk-at-prism.gatech.edu
Date: Wed, 27 Mar 1996 15:39:07 -0500 (EST)
Subject: Gold sources

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A note related to Richard Beanland's comment about using a flattened gold
coin as a sputter target to save money:

I use 1/4 oz Canadian Maple coins (purchased at a local coin dealer) as a
source of gold (99.99% pure) in my thermal vacuum evaporation system. I
cut them into small pie-shaped pieces using heavy duty wire cutters that
have been carefully cleaned and degreased. These pieces can be futher
subdivided prior to weighing them and placing them in the evaporator's
tungsten boat.

---------------------------------------------------------
Dr. Michael E. Knotts E-mail: ph281mk-at-prism.gatech.edu
Georgia Tech / School of Physics / Atlanta, GA 30332-0430
Tel: (404) 894-3422 FAX: (404) 894-9958




From: Damon Heer :      DLH-at-fei2.feico.com
Date: Wed, 27 Mar 1996 13:21:22 -0800
Subject: Focused Ion Beam Analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings list server friends. A recent request regarding FIB
information caught my eye and I forwarded it to a colleague who
works in our FIB systems division. His input is below. In an
effort to not clog up the list server with marketing or sales junk,
the response is mostly an offering of where to go to find further
information.

Cheers,
Damon

**************************************

Hello,

This indeed sounds like a FIB-related application. I am a Sales
Associate at FEI. As you may know, FEI specializes in the
development and production of FIB systems and their components. We
would be happy to answer any questions you have regarding this
technology.

In response to your friend's situation, we have an on-site lab that can
analyze any number of different types of samples, so please contact
me if you think I can be of assistance.

If you, your colleagues, or any others on the list server have any
additional FIB related inquiries, please feel free to e-mail me
directly, off line from this list server, at any time.

Brinker B. Gildersleeve
Sales Associate
FEI Company

bbg-at-feico.com



} Greetings,
}
} A colleague is needing compositional analysis from 10 nm layers. He is
} interested in doing focused ion beam analysis considering sample prep
} difficulties with TEM. Since I am not familiar with FIB, is this a viable
} technique for his problem? If it is, is there anyone out there willing to
} analyze his samples?
}
} Jeff Shield


FEI Company
7451 N.E. Evergreen Parkway
Hillsboro, OR 97124-5830

Phone (503) 640-7582
fax (503) 640-7509
email dlh-at-feico.com




From: Scott Holt :      102467.2752-at-CompuServe.COM
Date: 27 Mar 96 17:22:12 EST
Subject: EM Glue Suppliers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John Phelps wrote:

} We have been using Gatan "G-1" epoxy to prepare cross sections for TEM
} analysis. Does anyone know what this "G-1" epoxy is? More importantly, is
} this an epoxy that be purchased at, say a hardware store? Or perhaps some
} one knows of a substitute that works as well or possibly better that can be
} purchased locally.

} thanks in advance,
} John

} John Phelps
} NIST
} Materials Reliability Division
} 325 Broadway
} Boulder, CO 80303
} ph. 303-497-7570
} fax. 303-497-5030

John,

G-1 Epoxy appears to be the same as EPOXY TECHNOLOGY's 353ND epoxy. This
adhesive
was originally sold for fiber optic work. Anyway, we at BUEHLER sold it until
recently
when it's carcinogenic properties became better known to us. However, I believe
you
can still buy it directly from EPOXY TECHNOLOGY INC. in Billerica, MA (Phone:
1-800-227-2201).
Last time I bought some, I got a good sized can (without quantity written on the
label) estimated
at about 15-20oz. for $35.00. However, there are other EM suppliers who sell it
in smaller
containers for a somewhat reasonable price.

Good Luck.
Scott D. Holt
BUEHLER
41 Waukegan Rd.
Lake Bluff, IL 60044
(847)295-4546





From: slakmon-at-scottscientific.com (P. Slakmon)
Date: Wed, 27 Mar 1996 09:40:47 -0500
Subject: Conference & Workshop Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


S C O T T S C I E N T I F I C


On the behalf of The International Society of Molecular Morphology

Would like to take the opportunity to announce the

FOURTH INTERNATIONAL CONFERENCE & WORKSHOP
on
MOLECULAR MORPHOLOGY

June 3-4, 1996 - Conference
June 5 -6, 1996 - Workshop
in

Montreal, Canada

CONFERENCE FEATURING
Advances in Principles, Techniques and
Applications in Research and Diagnosis of:

- In Situ PCR
- In Situ Hybridization
- Immunohistochemistry
- Immunogold-Silver Staining
- Immunogold Electron Microscopy
- Microwave Immunohistochemistry
- Atomic Force Probe Microscopy
- Confocal Microscopy
- Antigen Retrieval
- Image Analysis

Call for Abstract Submission
DEADLINE: APRIL 15, 1996

The abstract should be typed single space on white paper within 9 x 7 inch
(23 x 18 cm) typed space. Total of two pages per abstract. Photographs and
references may be included. Please follow the style of CELL VISION, in
which the Proceedings will be published. Abstracts should be submitted in
duplicate.

THREE HANDS-ON WORKSHOPS (at the laboratories of Dept. of Anatomy,
University of Montreal)
1 In Situ PCR and In Situ Hybridization
2 Immunogold EM and Immunogold-Silver Staining
3 Microwave Fixation, Antigen Retrieval and Immunohistochemistry

June 2, 1996 (Sunday), 6.30-9.30 pm:
An optional preparation lecture on "Molecular Biology for the Uninitiated"

ORGANIZING COMMITTEE
CO-CHAIRMEN
Jiang Gu, M.D., Ph.D.
Deborah Research Institute
Browns Mills, New Jersey, USA

Moise Bendayan, Ph.D.
University of Montreal
Montreal, Canada

MEMBERS
Virginia Anderson, M.D.
Health Science Center at Brooklyn
State University of New York
Brooklyn, New York, USA

Gerhard Hacker, Ph.D.
Institute of Pathology
General Hospital, University of Salzburg
Salzburg, Austria

Lawrence DeBault, Ph.D.
Oklahoma University Health Center
Oklahoma City, Ok, USA

Shahla Masood, M.D.
University of Florida Health Science Center
Jackonsville, Florida, USA

Robert Day, Ph.D.
University of Montreal
Montreal, Canada

David Kersten, M.D.
Ealing
London, UK

----------------------------------------------------------------------------
-------------------------------------
ADVANCED REGISTRATION FORM
(please print and use photocopies for additional forms)

NAME ________________________________

PHONE _________________ FAX __________________

ADDRESS __________________________________________

_________________________________________

CHOICE OF WORKSHOP (circle one): 1 2 3
----------------------------------------------------------------------------
-------------------------------------
ADVANCED REGISTRATION FREE (for one of three):
- For the two-day conference $250 (US)
- For the two-day workshop $350 (US)

Make check payable to CELL VISION. A 15 % discount for members of the
"International Society of Molecular Morphology" and students (with proven
ID). A 15% discount will be reimbursed upon becoming a member of the
society before or at the conference.
Please book the hotel room directly by calling The Best Western Hotel in
Montreal, Canada (800) 361-3000 or Fax (514) 861-4089. You can obtain a
special discounted room rate at $79 (Canadian) per day (rate includes
breakfast) by identifying yourself as a participant of the conference/workshop.
Student dormitory available at University of Montreal (15-minute subway
transporation to conference location, 5 minute walk to workshop location) at
$35 (Canadian) per day by calling (514) 343-6531.

Send abstract, registration form, and registration fee to:
CELL VISION-JOURNAL OF ANALYTICAL MORPHOLOGY,
EDITORIAL OFFICE, DEBORAH RESEARCH INSTITUTE
1 Trenton Road, Browns Mills
NJ 08015-1799, USA
Phone: (609) 735-0477
Fax: (609) 735-0478


For further information please direct your inquiries by email to:
morphology-at-scottscientific.com
_______________________________________________
SCOTT SCIENTIFIC
P.O. Box 66552 Station Cavendish,
Montreal, Quebec, H4W 3J6, Canada

Telephone: 514-485-2309 Fax: 514-485-9931
Voice Mail: 514-888-6509

WWW Site: http://www.ScottScientific.com

E-Mail: slakmon-at-scottscientific.com
info-at-scottscientific.com
sales-at-scottscientific.com
admin-at-scottscientific.com
_________________________________________________





From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Wed, 27 Mar 1996 17:45:59 -0600
Subject: Atlas of prokaryotes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I need a reference to a through TEM/SEM atlas of ultrastructure of
prokaryotes and viruses. Or separate atlases for each.
Must be 1990 or later.
Thanks in advance for any info.
Phil

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: Bob Citron :      Bob_Citron-at-cc.chiron.com
Date: 3/26/96 8:54 AM
Subject: EMPA: anybody ever done hair?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Does anyone out there have any experience with EMPA of hair? Preparation,
mounting, etc? Someone here is interested in looking at heavy metals that
thru metabolism are concentrated in the hair.The levels will be low, so
the MDL will be a question, but I don't know how the hair will stand up to
an electron beam.

John


John Fournelle
Electron Microprobe Lab Internet:johnf-at-geology.wisc.edu
Dept of Geology & Geophysics Office: (608) 262-7964
University of Wisconsin Lab: (608) 265-4798
1215 West Dayton Street Fax: (608) 262-0693
Madison, WI 53706 Amateur radio: WA3BTA/9
http://geology.wisc.edu/~johnf/sx51.html

"The first rule of all intelligent tinkering is to save all the pieces."
Aldo Leopold








From: South Bay Technology :      73531.1344-at-CompuServe.COM
Date: 27 Mar 96 17:58:28 EST
Subject: Epoxy for TEM Sample Prep

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John-

The G1 is the same material we sell as "Golly G1" (we like to have some fun
here!).

The actual material is EpoTek 353ND. It is available from:

Epoxy Technology, Inc.
14 Fortune Drive
Billerica, MA 01821

TEL: 800-227-2201
FAX: 508-663-9782

We sell an 8 ounce kit of our Golly G1 for about $50. However, you can buy it
directly from Epoxy Technology for between $25-30. An 8 ounce kit will last you
a lifetime! I think Gatan sells it iin much smaller containers if that is
appealing (However, I don't think it costs any less).

Best regards-

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com
sbt-at-msa.microscopy.com

Manufacturers of Precision Sample Preparation Equipment and Supplies for
Metallography, Crystallography and Electron Microscopy.





From: Beverly E Maleeff
Date: 27 Mar 96 11:17:13 EDT
Subject: April PSM Meeting Notice

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9603271909.AA7252-at-pho903.sbphrd.com}
To: microscopy {microscopy-at-Sparc5.Microscopy.Com}
{Beverly_E_Maleeff-at-sbphrd.com}

Philadelphia Society for Microscopy Meeting Notice
April 1996


DATE: Wednesday, April 10, 1996

PLACE: Laboratory for the Research of Science and Materials
(LRSM) Building, 33rd and Walnut Street, Philadelphia, PA.
Parking is available behind the LRSM Building after 5:00 PM.


TIME: 5:30 - 7:30 PM Social hour, hosted by our meeting sponsor.
Buffet dinner service and informal seating during this time

7:30 PM Speaker:

Photooxidation of Fluorescent Markers:
A Bridge Between Light and Electron Microscopic Cytochemistry

Dr. Giuseppe G. Pietra
Professor of Pathology
Hospital of the University of Pennsylvania
Philadelphia, PA

Abstract:
Fluorescent labeling techniques have become very popular in many areas of
pathology and cell biology. However, a limitation of fluorescent labeling is
the relative low resolution of the optical microscope, even using laser
scanning confocal microscopy. Excitation of a fluorescent dye in the presence
of 3-3-diaminobenzidine (DAB) oxidizes DAB to an insoluble electron-dense
product that can be readily localized by electron microscopy. The application
of this method to biological problems will be illustrated in a correlative
confocal and electron microscopic study of experimental pulmonary edema.
Advantages and limitations of this technique will be discussed.




DINNER: With a Mexican Flair!!

COST: Members $12.00 Student members $6.00 Non-members $15.00

MENU:
Margaritas
Mexican beer
Salsa, chips, pretzels, etc.

Vegetarian chili
Taco shells, crispy or soft (matzoh will be available)
Sauteed ground beef or chicken
Shredded lettuce
Grated cheese
Diced tomatoes
Sour cream
Guacamole
Refried beans
Tortilla chips

Fresh fruit salad

Coffee, decaf or tea


Reservations will be taken by Ms. Pat Overend at the University of
Pennsylvania, 215/898-8337. Deadline for reservations will be Friday, April
5. If you have any questions regarding the meeting please feel free to contact
Rollin Lakis at 215/898-2013 or lakis-at-sol1.lrsm.upenn.edu.
Cancellations must be received by Ms. Overend no later than 5:00 PM, April 5,
1996.






From: David_Gantt-at-GSVMS2.CC.GASOU.EDU (DAVID G. GANTT)
Date: Wed, 27 Mar 1996 20:07:39 -0400
Subject: Tooth buds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01530501ad7f888249c1-at-[141.165.35.119]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I am interested in determining the best procedure to study tooth buds with
the SEM. Any Suggestion!

Dr. David G. Gantt Phone: 1-912-681-5964
Dept. of Biology Fax: 1-912-681-0845
Landrum Box 8042 e-mail: david_gantt-at-gsvms2.cc.gasou.edu
Georgia Southern University
Statesboro, Georgia 30460-8042








From: Sarka Lhotak :      lhotaks-at-fhs.csu.mcmaster.ca
Date: Wed, 27 Mar 1996 19:55:25 -0500 (EST)
Subject: Re: Embedding w/ LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Just a quick note on LR White embedding: We occasionally use flat
embedding with LR White into polyethylene (I believe) molds and cover
the whole thing with Saran wrap. It is a bit messy but it works. I believe
you could overfill the tube sligthly and cover with Saran wrap too.
Cheers,

Sarka Lhotak
EM Facility, McMaster University
Hamilton, Ontario, Canada
lhotaks-at-fhs.mcmaster.ca




From: James S MArtin :      James.S.Martin-at-williams.edu
Date: Wed, 27 Mar 1996 22:24:33 -0500 (EST)
Subject: re: EMPA of hair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential
Microscopy ListServer {Microscopy-at-Sparc5.Microscopy.Com}
Message-Id: {960327154237.718-at-cliff.ml.wpafb.af.mil.0}

Several years ago, 1991 I think, a forensic scientist from Scotland Yard was
the dinner speaker at Inter-Micro in Chicago. One of the cases he
described involved the poisoning of a woman from the Middle East. He
detected arsenic along a strand of the woman's hair using SEM-EDS. By
correlating the intervals at which arsenic was detected with the growth
rate of the woman's hair and her schedule over the preceding months, he
demonstrated that she was intentially and repeatedly poisoned in the same
city. I recall that this led to an arrest. If you need more info, let
me know.

James Martin
Williamstown Art Conservation Center
Williamstown, MA




From: John M. Libert :      jlibert-at-cpcug.org
Date: Wed, 27 Mar 1996 20:55:25 -0500
Subject: Re: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {3159F18D.342A-at-cpcug.org}

scott_l-at-unin1.unorth.ac.za wrote:
}
} Are there IBM compatible PC based programmes able to assist with doing
} stereology from TEM and LM micrographs. What are the hardware require-
} ments?
}
} Thank you
} Leon Scott
} SAYes. Assuming your images are in TIFF or other common format, the
Bioquant software from R&M Biometrics has an excellent stereology toolkit
add-on module to the True Color Windows image analysis and topographic
morphometry software package. R&M likes to provide the PC integrated with
the software and the high-end frame grabber that they use, but the system
can be installed on most 486 or Pentium computers able to run Windows.
Generally, 16MB RAM and large disk capacity (1 GByte) is recommended for
image work. If you wish, I can help you locate the BioQuant dealer in
your area. Or, better you can call R&M's office in Nashville, TN at
1-800-221-0549. They will be happy to send you literatue and even to
arrange for a demonstration for you.

Regards,

John Libert




From: John M. Libert :      jlibert-at-cpcug.org
Date: Wed, 27 Mar 1996 20:55:25 -0500
Subject: Re: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {3159F18D.342A-at-cpcug.org}

scott_l-at-unin1.unorth.ac.za wrote:
}
} Are there IBM compatible PC based programmes able to assist with doing
} stereology from TEM and LM micrographs. What are the hardware require-
} ments?
}
} Thank you
} Leon Scott
} SAYes. Assuming your images are in TIFF or other common format, the
Bioquant software from R&M Biometrics has an excellent stereology toolkit
add-on module to the True Color Windows image analysis and topographic
morphometry software package. R&M likes to provide the PC integrated with
the software and the high-end frame grabber that they use, but the system
can be installed on most 486 or Pentium computers able to run Windows.
Generally, 16MB RAM and large disk capacity (1 GByte) is recommended for
image work. If you wish, I can help you locate the BioQuant dealer in
your area. Or, better you can call R&M's office in Nashville, TN at
1-800-221-0549. They will be happy to send you literatue and even to
arrange for a demonstration for you.

Regards,

John Libert




From: slakmon-at-scottscientific.com (P. Slakmon)
Date: Wed, 27 Mar 1996 19:44:00 -0500
Subject: Conference & Workshop Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


S C O T T S C I E N T I F I C


On the behalf of The International Society of Molecular Morphology

Would like to take the opportunity to announce the

FOURTH INTERNATIONAL CONFERENCE & WORKSHOP
on
MOLECULAR MORPHOLOGY

June 3-4, 1996 - Conference
June 5 -6, 1996 - Workshop
in

Montreal, Canada

CONFERENCE FEATURING
Advances in Principles, Techniques and
Applications in Research and Diagnosis of:

- In Situ PCR
- In Situ Hybridization
- Immunohistochemistry
- Immunogold-Silver Staining
- Immunogold Electron Microscopy
- Microwave Immunohistochemistry
- Atomic Force Probe Microscopy
- Confocal Microscopy
- Antigen Retrieval
- Image Analysis

Call for Abstract Submission
DEADLINE: APRIL 15, 1996

The abstract should be typed single space on white paper within 9 x 7 inch
(23 x 18 cm) typed space. Total of two pages per abstract. Photographs and
references may be included. Please follow the style of CELL VISION, in
which the Proceedings will be published. Abstracts should be submitted in
duplicate.

THREE HANDS-ON WORKSHOPS (at the laboratories of Dept. of Anatomy,
University of Montreal)
1 In Situ PCR and In Situ Hybridization
2 Immunogold EM and Immunogold-Silver Staining
3 Microwave Fixation, Antigen Retrieval and Immunohistochemistry

June 2, 1996 (Sunday), 6.30-9.30 pm:
An optional preparation lecture on "Molecular Biology for the Uninitiated"

ORGANIZING COMMITTEE
CO-CHAIRMEN
Jiang Gu, M.D., Ph.D.
Deborah Research Institute
Browns Mills, New Jersey, USA

Moise Bendayan, Ph.D.
University of Montreal
Montreal, Canada

MEMBERS
Virginia Anderson, M.D.
Health Science Center at Brooklyn
State University of New York
Brooklyn, New York, USA

Gerhard Hacker, Ph.D.
Institute of Pathology
General Hospital, University of Salzburg
Salzburg, Austria

Lawrence DeBault, Ph.D.
Oklahoma University Health Center
Oklahoma City, Ok, USA

Shahla Masood, M.D.
University of Florida Health Science Center
Jackonsville, Florida, USA

Robert Day, Ph.D.
University of Montreal
Montreal, Canada

David Kersten, M.D.
Ealing
London, UK

----------------------------------------------------------------------------
-------------------------------------
ADVANCED REGISTRATION FORM
(please print and use photocopies for additional forms)

NAME ________________________________

PHONE _________________ FAX __________________

ADDRESS __________________________________________

_________________________________________

CHOICE OF WORKSHOP (circle one): 1 2 3
----------------------------------------------------------------------------
-------------------------------------
ADVANCED REGISTRATION FREE (for one of three):
- For the two-day conference $250 (US)
- For the two-day workshop $350 (US)

Make check payable to CELL VISION. A 15 % discount for members of the
"International Society of Molecular Morphology" and students (with proven
ID). A 15% discount will be reimbursed upon becoming a member of the
society before or at the conference.
Please book the hotel room directly by calling The Best Western Hotel in
Montreal, Canada (800) 361-3000 or Fax (514) 861-4089. You can obtain a
special discounted room rate at $79 (Canadian) per day (rate includes
breakfast) by identifying yourself as a participant of the conference/workshop.
Student dormitory available at University of Montreal (15-minute subway
transporation to conference location, 5 minute walk to workshop location) at
$35 (Canadian) per day by calling (514) 343-6531.

Send abstract, registration form, and registration fee to:
CELL VISION-JOURNAL OF ANALYTICAL MORPHOLOGY,
EDITORIAL OFFICE, DEBORAH RESEARCH INSTITUTE
1 Trenton Road, Browns Mills
NJ 08015-1799, USA
Phone: (609) 735-0477
Fax: (609) 735-0478


For further information please direct your inquiries by email to:
morphology-at-scottscientific.com
_______________________________________________
SCOTT SCIENTIFIC
P.O. Box 66552 Station Cavendish,
Montreal, Quebec, H4W 3J6, Canada

Telephone: 514-485-2309 Fax: 514-485-9931
Voice Mail: 514-888-6509

WWW Site: http://www.ScottScientific.com

E-Mail: slakmon-at-scottscientific.com
info-at-scottscientific.com
sales-at-scottscientific.com
admin-at-scottscientific.com
_________________________________________________





From: slakmon-at-scottscientific.com (P. Slakmon)
Date: Wed, 27 Mar 1996 19:34:37 -0500
Subject: Conference & Workshop Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


S C O T T S C I E N T I F I C


On the behalf of The International Society of Molecular Morphology

Would like to take the opportunity to announce the

FOURTH INTERNATIONAL CONFERENCE & WORKSHOP
on
MOLECULAR MORPHOLOGY

June 3-4, 1996 - Conference
June 5 -6, 1996 - Workshop
in

Montreal, Canada

CONFERENCE FEATURING
Advances in Principles, Techniques and
Applications in Research and Diagnosis of:

- In Situ PCR
- In Situ Hybridization
- Immunohistochemistry
- Immunogold-Silver Staining
- Immunogold Electron Microscopy
- Microwave Immunohistochemistry
- Atomic Force Probe Microscopy
- Confocal Microscopy
- Antigen Retrieval
- Image Analysis

Call for Abstract Submission
DEADLINE: APRIL 15, 1996

The abstract should be typed single space on white paper within 9 x 7 inch
(23 x 18 cm) typed space. Total of two pages per abstract. Photographs and
references may be included. Please follow the style of CELL VISION, in
which the Proceedings will be published. Abstracts should be submitted in
duplicate.

THREE HANDS-ON WORKSHOPS (at the laboratories of Dept. of Anatomy,
University of Montreal)
1 In Situ PCR and In Situ Hybridization
2 Immunogold EM and Immunogold-Silver Staining
3 Microwave Fixation, Antigen Retrieval and Immunohistochemistry

June 2, 1996 (Sunday), 6.30-9.30 pm:
An optional preparation lecture on "Molecular Biology for the Uninitiated"

ORGANIZING COMMITTEE
CO-CHAIRMEN
Jiang Gu, M.D., Ph.D.
Deborah Research Institute
Browns Mills, New Jersey, USA

Moise Bendayan, Ph.D.
University of Montreal
Montreal, Canada

MEMBERS
Virginia Anderson, M.D.
Health Science Center at Brooklyn
State University of New York
Brooklyn, New York, USA

Gerhard Hacker, Ph.D.
Institute of Pathology
General Hospital, University of Salzburg
Salzburg, Austria

Lawrence DeBault, Ph.D.
Oklahoma University Health Center
Oklahoma City, Ok, USA

Shahla Masood, M.D.
University of Florida Health Science Center
Jackonsville, Florida, USA

Robert Day, Ph.D.
University of Montreal
Montreal, Canada

David Kersten, M.D.
Ealing
London, UK

----------------------------------------------------------------------------
-------------------------------------
ADVANCED REGISTRATION FORM
(please print and use photocopies for additional forms)

NAME ________________________________

PHONE _________________ FAX __________________

ADDRESS __________________________________________

_________________________________________

CHOICE OF WORKSHOP (circle one): 1 2 3
----------------------------------------------------------------------------
-------------------------------------
ADVANCED REGISTRATION FREE (for one of three):
- For the two-day conference $250 (US)
- For the two-day workshop $350 (US)

Make check payable to CELL VISION. A 15 % discount for members of the
"International Society of Molecular Morphology" and students (with proven
ID). A 15% discount will be reimbursed upon becoming a member of the
society before or at the conference.
Please book the hotel room directly by calling The Best Western Hotel in
Montreal, Canada (800) 361-3000 or Fax (514) 861-4089. You can obtain a
special discounted room rate at $79 (Canadian) per day (rate includes
breakfast) by identifying yourself as a participant of the conference/workshop.
Student dormitory available at University of Montreal (15-minute subway
transporation to conference location, 5 minute walk to workshop location) at
$35 (Canadian) per day by calling (514) 343-6531.

Send abstract, registration form, and registration fee to:
CELL VISION-JOURNAL OF ANALYTICAL MORPHOLOGY,
EDITORIAL OFFICE, DEBORAH RESEARCH INSTITUTE
1 Trenton Road, Browns Mills
NJ 08015-1799, USA
Phone: (609) 735-0477
Fax: (609) 735-0478


For further information please direct your inquiries by email to:
morphology-at-scottscientific.com
_______________________________________________
SCOTT SCIENTIFIC
P.O. Box 66552 Station Cavendish,
Montreal, Quebec, H4W 3J6, Canada

Telephone: 514-485-2309 Fax: 514-485-9931
Voice Mail: 514-888-6509

WWW Site: http://www.ScottScientific.com

E-Mail: slakmon-at-scottscientific.com
info-at-scottscientific.com
sales-at-scottscientific.com
admin-at-scottscientific.com
_________________________________________________





From: Bo Johansen :      BOJ-at-bot.ku.dk
Date: Thu, 28 Mar 1996 11:06:08 GMT+0100
Subject: Elmers glue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello - is there anybody out there who can tell me where
to obtain 'Elmers glue'.

Bo


_____________________________________________________________________
Bo Johansen E-Mail: BoJ-at-bot.ku.dk
Botanical Institute Vioce: +45 3532 2157
Gothersgade 140 FAX: +45 3313 9104
DK-1123 Copenhagen K, Denmark http://www.bot.ku.dk/www/staff/boj.htm
---------------------------------------------------------------------






From: James S MArtin :      James.S.Martin-at-williams.edu
Date: Thu, 28 Mar 1996 08:27:36 -0500 (EST)
Subject: Re: Elmers glue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Elmer's Glue is manufactured by Borden, Inc., Columbus 43215 (USA).
Borden makes at least two different types of Elmer's (registered
tradename) glue -- a white glue composed primarily of an aqueous
dispersion of poly(vinyl acetate), and a yellow carpenter's wood glue
composed primarily of an aliphatic resin.

James Martin
Williamstown Art Conservation Center


On Thu, 28 Mar 1996, Bo Johansen wrote:

} Hello - is there anybody out there who can tell me where
} to obtain 'Elmers glue'.
}
} Bo
}
}
} _____________________________________________________________________
} Bo Johansen E-Mail: BoJ-at-bot.ku.dk
} Botanical Institute Vioce: +45 3532 2157
} Gothersgade 140 FAX: +45 3313 9104
} DK-1123 Copenhagen K, Denmark http://www.bot.ku.dk/www/staff/boj.htm
} ---------------------------------------------------------------------
}
}
}




From: colijn.1-at-osu.edu (Henk Colijn)
Date: Thu, 28 Mar 1996 09:01:42 -0500 (EST)
Subject: Post Doc position available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Post-Poctoral Research Associate
Center for Industrial Sensors and Measurements
The Ohio State University

A post-doctoral research position is presently available in the Department of
Materials Science and Engineering in the area of TEM characterization of
ceramic-based materials. Specifically this position is part of a new NSF
Center for Industrial Sensors and Measurements (CISM). Key components of this
work will be analytical (EELS and EDS) and high resolution TEM investigation of
interfaces in ceramic systems such as those based on TiO2 and Zr2O3. The state-
of-the-art microscopy facilities available in the Department of Materials
Science and Engineering will be utilized for this program. The position will
also involve close interaction with a number of industrial partners who are
collaborating in this effort to develop improved sensors for harsh industrial
environments.

Applicants may send a cirriculum vitae by mail or E-mail to:

Professor Michael J. Mills
Department of Materials Science and Engineering
The Ohio State University
477 Watts Hall
2041 College Road
Columbus, OH 43210-1178

Tel: (614) 292 - 7514
Fax: (614) 292 - 1537

E-mail: mills.108-at-osu.edu

Henk Colijn colijn.1-at-osu.edu OSU Campus Electron Optics Facility
-------------------------------------------------------------------
An optimist believes that we live in the best of all possible worlds.
A pessimist fears this is true.






From: RMacKay :      RMACKAY-at-AC.DAL.CA
Date: Thu, 28 Mar 1996 10:35:05 +0000
Subject: Re: EMPA: anybody ever done hair?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603281440.KAA09558-at-Snoopy.UCIS.Dal.Ca}
Comments: Authenticated sender is {RMACKAY-at-ac.dal.ca}

John,

R. A. Smith did some work in this area a few years ago. See
" A Method to Distinguish between Arsenic in and on human
Hair ", Environmental Research 12, 171-173 (1976).

Bob MacKay
Robert MacKay
Department of Earth Sciences
Dalhousie University
Halifax, Nova Scotia, Canada
B3H 3J5
Tel: 902 494-7087
e-mail rmackay-at-ac.dal.ca




From: Len Adleman :      adleman-at-pollux.usc.edu
Date: Thu, 28 Mar 1996 10:34:04 -0800 (PST)
Subject: Please unsubscibe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9603281448.AA04445-at-crdems.ge.com}

I would like to unsubscribe:
adleman-at-cs.usc.edu
The volume was too great for my modest needs - thnaks




From: South Bay Technology :      73531.1344-at-CompuServe.COM
Date: 28 Mar 96 12:21:21 EST
Subject: TEM: Copy of Ultramicroscopy paper

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friends-

Can anyone out there supply me with a copy of the following paper?

"Applications of an Ionless Tripod Polisher to Preparation of YBCO
Superconducting Multilayer and Bulk Ceramic Thin Films" J. Ayache & P. H.
Alboreole.

Ultramicroscopy, 60(2), September 1995

Thank you!

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com
sbt-at-msa.microscopy.com

Manufacturers of Precision Sample Preparation Equipment and Supplies for
Metallography, Crystallography and Electron Microscopy.





From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 28 Mar 1996 09:20:51 -0400
Subject: RE- CalibStds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1384117519.71286-at-mse.engin.umich.edu}

Subject: Time: 9:14 AM
OFFICE MEMO RE: CalibStds Date: 3/28/96

Bob:
I am afraid that you will find that logical arguments, based on physical
reality, such as you offer regarding the stability of standards for
calibrating optical microscopes, don't wash with the bureaucracies involved
in promulgating and enforcing standardization and safety regulations. Once a
regulation is written to take care of one type of problem, it is applied
rigorously in all other situations, no matter how ridiculous. I recently
heard of a incident in which a safety inspector required a label listing
manufacturer, composition, safety hazards, etc. on a wash bottle that was
clearly marked as containing distilled water.
Wil Bigelow (bigelow-at-umich.edu)





From: Hand-at-nso1.uchc.edu (Hand,Arthur)
Date: Thu, 28 Mar 1996 09:52:35 -0500
Subject: EM User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


--=====================_19263778==_
Content-Type: text/plain; charset="US-ASCII"

Since there have been a large number of requests for the data on user fees, I
have decided to post the information on the listserver as attachments to
this message. They were created in MS Word 6; one attachment is in Mac
format; the other is in PC format. If you have problems opening them, let me
know and I will fax the data to you as soon as possible.

Arthur R. Hand
Central EM Facility
UConn Health
Ctr

--=====================_19263778==_
Content-Type: application/octet-stream; name="Fee Comparison-Mac"
Content-Transfer-Encoding: x-uuencode

begin 666 Fee Comparison-Mac
MT,\1X*&Q&N$`````````````````````.P`#`/[_"0`&```````````````!
M`````0``````````$````-at-````$```#^____``````````#_____________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M_______________________]____#-at-```/[___\/````!0````8````'````
M"`````D````*````"P````P````-````$````/[____^____$0```!(````3
M````%````!4````6````%P```!-at-````9````&-at-```!L```` {````'0```!X`
M```?````(````"$````B````(P```"0````E````)-at-```" {```#^________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M_____________________________________________P!2````````````
M````````````````````````````````````````````````````````````
M```````````"``4`__________\#``````D"``````#`````````1-at-``````
M`````````(`3BXN+'+L!`P```(`"`````````0!#`&\`;0!P`$\`8-at-!J````
M````````````````````````````````````````````````````````````
M`!(``-at-'_______________\`````````````````````````````````````
M````````````````7`````````!7`&\` {-at-!D`$0`;P!C`'4`;0!E`&X`=```
M````````````````````````````````````````````````````&-at-`"`?__
M__\$````_____P``````````````````````````````````````````````
M``0```#L0-at-```````$\`8-at-!J`&4`8P!T`%``;P!O`&P`````````````````
M```````````````````````````````````````````6``$!`0````(```#_
M____``````````````````````````"`YEF*BQR[`8#F68J+'+L!````````
M`````````0```/[___\#````!`````4````&````!P````-at-````)````_O__
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M______________________________\!`/[_`-at-`!`/____\`"0(``````,``
M``````!&'````$UI8W)O {V]F="!7;W)D(#8N,"!$;V-U;65N=`#^____3D(V
M5Q````!7;W)D+D1O8W5M96YT+C8````````[``/__-at-`)``8`````````````
M`````0````$````````0`/[_```#"-at-$```````````````````````$```#R
MGX7-at-3_D0:*N1"``K)[/9,````*X!```/````!P```)-at-````"````X`````0`
M```#`0``"````!H!```!````,0$```P````W`0``"P```$,!```-````3P$`
M``\```!;`0``$````&,!```*````:P$``!(```!W`0``#-at-```)0!```)````
MG`$``!,```"F`0``________________________________'-at-```$````!-
M86-I;G1O {V-at--at-2$0Z36EC {F]S;V9T($]F9FEC93I-:6-R;W-O9G0-at-5V]R9"`V
M.E1E;7!L871E {SI.;W)M86P`'-at-```!L```!#3TU005))4T].($]&($5-(%53
M15(-at-1D5%4P`} ````#P```$%R=&AU {B!2+B!(86YD`!X````/````07)T:'5R
M(%(N($AA;F0``-at-```!`G0`````!"FH*+'+L!0`````"`=M/RS. {!0-RE:`!/
MP`D$`````&4```````$```$``0`#``#C$0``[$(`````````````````````
M``#C#-at-``````````````````````````````````````````````/-at-``;```
M```^``!L````;#X```````!L/-at-```````&P^````````;#X```````!L/-at-``
M%````+8^````````MCX```````"V/-at-```````+8^````````MCX```````"V
M/-at-``"-at-```,`^``!\````MCX````````-0-at-``/0```#P_````````/#\`````
M```\/P```````#P_````````/#\````````\/P```````#P_````````/#\`
M``````#E00```-at-```.=!````````YT$```````#G00```````.=!````````
MYT$```````#G00``)-at-```$I"``!8````HD(``$H````-0-at-``````````````
M``````````````!L/-at-```````#P_```````````*``L``0`4`#P_````````
M/#\`````````````````````````````/#\````````\/P````````U"````
M````/#\```````!L/-at-```````&P^````````/#\`````````````````````
M````````/#\````````\/P```````#P_````````/#\````````\/P``````
M`&P^````````/#\```````!L/-at-```````#P_````````Y4$`````````````
M````````````````-at-#X``!0```"4/-at-``(-at-```&P^````````;#X```````!L
M/-at-```````&P^````````/#\```````#E00```````#P_``"I`-at-``/#\`````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````````````````````````U#
M3TU005))4T].($]&($5-(%5315(-at-1D5%4R!&3U(-at-4T5,14-4140-at-4T525DE#
M15,L($UA {F-H(#$Y.38-#0=#96YT {F%L($5-($9A8VEL:71Y+"!50V]N;-at-=5
M($UI8V-at--at-4V-H(&]F($1E;G1I {W1R} 0=5($9L;W)I9&$-14T-at-0V]R90=5($AA
M=V%I:2P-at-4&%C($)I;VUE9"!297,-at-0W1R!P=414TL(%-%32!U {V4L('!E {B!H
M {-at- {'!P {'!R`-at-(%5N:79E {G-I='D'-3`N,#`',S`N,#`',C4N,#`',S(N,#`M
M,S8N,#`'!R`-at-($EN9'5S=')Y!S4P+C`P!S$P,"XP,` {'!P=5;'1R86UI8W)O
M=&]M92!U {V4L('!E {B!H {-at- {Q,"XP,` {'!S8N,#`'!U-A;7!L92!P {F5P87)A
M=&EO;-at- {'!P {'!R`-at-($-O;7!L971E(%1%30 {Q,3`N,#`',34P+C`P!S {U+C`P
M!P {'("`-at-("`-at-($9I} &%T:6]N(&%N9"!E;6)E9&1I;F {',S`N,#`',3`N,#`'
M,C4N,#`'!P {-at-("`-at-("`-at-5&AI;B!S96-T:6]N:6YG!S4P+C`P!S$R+C4P!S(U
M+C`P!P {'("`-at-("`-at-($=R:60-at- {W1A:6YI;F {'*&EN8VQU9&5D(&EN(&%B;W9E
M*0 {Q,"XP,` {'!P {-at-("!#;VUP;&5T92!314TL(&)I;VQO9VEC86P',3(P+C`P
M!P {W-2XP,` {'!R`-at-("`-at-("!3 {&5C:6UE;B!P {F5P(&]N;'D'-3`N,#`'-#`N
M,#`'!P {'("`-at-0V]M {&QE=&4-at-4T5-+"!N;VXM8FEO;` {X,"XP,` {'!P {'("`-at-
M("`-at-(%-P96-I;65N('!R97`-at-;VYL} 0 {S,"XP,` {R,"XP,` {'!P=.96=A=&EV
M92!S=&%I;B`H8V]M {&QE=&4I!S4P+C`P!P {T,"XP,` {'!TYE9V%T:79E('-T
M86EN("AP {F5P(&]N;'DI!S(P+C`P!P {'!P=& {F5E} F4-at-9G)A8W1U {F4-at-*'!R
M97`-at-;VYL} 2D',3`P+C`P!P {'!P=& {F5E} F4-at-9G)A8W1U {F4-at-:6YS='(-at-=7-E
M+VAR!P {'!P {'14T-at-=&5C:&YI8V%L(&AE;'`O:'('!P {',C4N,#`'!PT'3VAI
M;R!3=&%T92!5($5L96-T($]P=&EC {R!&86,'2DA5($5A {G1H("8-at-4&QA;F5T
M(%-C:0=2=71G97)S($UA {FEN92`F($-O87-T86P-at-4V-I!U4-at-36ES {V]U {FD-at-
M#5)O;&QA!P=414TL(%-%32!U {V4L('!E {B!H {-at- {'!P {'!R`-at-(%5N:79E {G-I
M='D',34N,#`M,S4N,#`',S4N,#`',S4N,#`'-S4N,#`'!R`-at-($EN9'5S=')Y
M!SDP+C`P+3(P,"XP,` {'!P {'56QT {F%M:6-R;W1O;64-at-=7-E+"!P97(-at-:'('
M!P {'!P=386UP;&4-at- {')E {&%R871I;VX'!P {'!P {-at-("!#;VUP;&5T92!414T'
M!P {'!P {-at-("`-at-("`-at-1FEX871I;VX-at-86YD(&5M8F5D9&EN9P {'!P {'!R`-at-("`-at-
M("!4:&EN('-E8W1I;VYI;F {'!P {'!P {-at-("`-at-("`-at-1W)I9"!S=&%I;FEN9P {'
M!P {'!R`-at-($-O;7!L971E(%-%32P-at-8FEO;&]G:6-A;` {'!P {'!R`-at-("`-at-("!3
M {&5C:6UE;B!P {F5P(&]N;'D'!P {'!P {-at-("!#;VUP;&5T92!314TL(&YO;BUB
M:6]L!P {'!P {'("`-at-("`-at-(%-P96-I;65N('!R97`-at-;VYL} 0 {'!P {'!TYE9V%T
M:79E('-T86EN("AC;VUP;&5T92D'!P {'!P=.96=A=&EV92!S=&%I;B`H {')E
M {"!O;FQY*0 {'!P {'!T9R965Z92!F {F%C='5R92`H {')E {"!O;FQY*0 {'!P {'
M!T9R965Z92!F {F%C='5R92!I;G-T {B!U {V4O:'('!P {'!P=%32!T96-H;FEC
M86P-at-:&5L {"]H {-at- {R-2XP,` {',C`N,#`',C4N,#`'!PT-!U4-at-36%R} 6QA;F0-at-
M0F%L=&EM;W)E($-T} 0=5($UA {W,-at-06UH97)S="!":6]L;V=Y!U-O($EL;&EN
M;VES($-T {B!F;W(-at-14T'0V]R;F5L;"!-870-at-4V-I($9A8VEL:71Y!P=414TL
M(%-%32!U {V4L('!E {B!H {-at- {'!P {'!R`-at-(%5N:79E {G-I='D',S$N,C4',34N
M,#`',3(N-3`',3 {N,#`M,S,N,#`'!R`-at-($EN9'5S=')Y!P {'!P {'56QT {F%M
M:6-R;W1O;64-at-=7-E+"!P97(-at-:'(',3(N-3`'!S$P+C`P!P {'4V%M {&QE('!R
M97!A {F%T:6]N!P {'!P {'("`-at-0V]M {&QE=&4-at-5$5-!S(U,"XP,` {'.34N,#`'
M!P {-at-("`-at-("`-at-1FEX871I;VX-at-86YD(&5M8F5D9&EN9P {T,"XP,` {Q,"XP,` {T
M,"XP,` {'!R`-at-("`-at-("!4:&EN('-E8W1I;VYI;F {',34N,#`',3`N,#`',C`N
M,#`'!P {-at-("`-at-("`-at-1W)I9"!S=&%I;FEN9P {'!S$U+C`P!P {'("`-at-0V]M {&QE
M=&4-at-4T5-+"!B:6]L;V=I8V%L!S(W-2XP,` {'!P {'("`-at-("`-at-(%-P96-I;65N
M('!R97`-at-;VYL} 0 {W-2XP,` {U-2XP,` {T,"XP,` {'!R`-at-($-O;7!L971E(%-%
M32P-at-;F]N+6)I;VP'!P {'!P {-at-("`-at-("`-at-4W!E8VEM96X-at- {')E {"!O;FQY!P {'
M,C`N,#`'!P=.96=A=&EV92!S=&%I;B`H8V]M {&QE=&4I!P {'!P {'3F5G871I
M=F4-at- {W1A:6X-at-*'!R97`-at-;VYL} 2D'!S$P+C`P!P {'!T9R965Z92!F {F%C='5R
M92`H {')E {"!O;FQY*0 {'-#`N,#`'!P {'1G)E97IE(&9R86-T=7)E(&EN {W1R
M('5S92]H {-at- {',C`N,#`'!P {'14T-at-=&5C:&YI8V%L(&AE;'`O:'('!P {',S(N
M,#`'!PT'26]W82!3=&%T92!"97-S97D-at-36EC($9A8P=,;WEO;&$-at-52!-0PU#
M96QL($)I;VP'52!/:VQA:&]M82P-3F]R;6%N!U4-at-0V%L:68L($1A=FES#49A
M8R!!9'8-at-26YS='('!U1%32P-at-4T5-('5S92P-at- {&5R(&AR!P {'!P {'("`-at-56YI
M=F5R {VET} 0 {R-RXU,` {Q,"XP,` {R-"XU,"TS-"XP,` {Q-2XP,` {'("`-at-26YD
M=7-T {GD'!P {',3 {U+C`P!P=5;'1R86UI8W)O=&]M92!U {V4L('!E {B!H {-at- {V
M+C8P!S$P+C`P!RAC {GEO+V1A} 2D-at-(#$U+C`P!P {'4V%M {&QE('!R97!A {F%T
M:6]N!P {'!P {'("`-at-0V]M {&QE=&4-at-5$5-!P {Q,C(N-3`'!P {'("`-at-("`-at-($9I
M} &%T:6]N(&%N9"!E;6)E9&1I;F {',38N-3`'!P {'!R`-at-("`-at-("!4:&EN('-E
M8W1I;VYI;F {'!S$U+C`P!P {'!R`-at-("`-at-("!' {FED('-T86EN:6YG!P {'!P {'
M("`-at-0V]M {&QE=&4-at-4T5-+"!B:6]L;V=I8V%L!P {Y,"XP,` {'!P {-at-("`-at-("`-at-
M4W!E8VEM96X-at- {')E {"!O;FQY!RA&:7-at-F0U!$*2`-at-,38N-3`'!P {'!R`-at-($-O
M;7!L971E(%-%32P-at-;F]N+6)I;VP'!P {'!P {-at-("`-at-("`-at-4W!E8VEM96X-at- {')E
M {"!O;FQY!P {'!P {'3F5G871I=F4-at- {W1A:6X-at-*&-O;7!L971E*0 {'!P {'!TYE
M9V%T:79E('-T86EN("AP {F5P(&]N;'DI!S8N-C`'!P {'!T9R965Z92!F {F%C
M='5R92`H {')E {"!O;FQY*0 {'!P {'!T9R965Z92!F {F%C='5R92!I;G-T {B!U
M {V4O:'('!P {H {&5R(&1A} 2D-at-("`-at-,3`N,#`'!P=%32!T96-H;FEC86P-at-:&5L
M {"]H {-at- {Q-BXU,` {H4V-O {&4I("`-at-("`Q,"XP,` {'!P {-#0=%4$9,+4-)344-
M3&%U {V%N;F4L(%-W:71Z!U1R96YT(%4-3VYT87)I;RP-at-0V%N!U4-at-;V8-at-5'5R
M:W4-1FEN;&%N9`=5;FEV($UE9"!38V-at--1&5B {F5C96XL($AU;F {'!U1%32P-at-
M4T5-('5S92P-at- {&5R(&AR!P {'!P {'("`-at-56YI=F5R {VET} 0 {R,"XP,` {Q,"XP
M,"TQ-2XP,` {Q-RXU,"TT-"XU,` {'!R`-at-($EN9'5S=')Y!RHS,#`N,#`M-3`P
M+C`P!P {'!P=5;'1R86UI8W)O=&]M92!U {V4L('!E {B!H {-at- {'!P {'!U-A;7!L
M92!P {F5P87)A=&EO;-at- {'!P {'!R`-at-($-O;7!L971E(%1%30 {'!P {H=R]O('!H
M;W1O*2`-at-("`-at--RXP,` {'("`-at-("`-at-($9I} &%T:6]N(&%N9"!E;6)E9&1I;F {'
M!P {S+C(U!P {'("`-at-("`-at-(%1H:6X-at- {V5C=&EO;FEN9P {'!RAP97(-at-9W)I9"D-at-
M+C {U!P {'("`-at-("`-at-($=R:60-at- {W1A:6YI;F {'!P {'!P {-at-("!#;VUP;&5T92!3
M14TL(&)I;VQO9VEC86P'!P {'!P {-at-("`-at-("`-at-4W!E8VEM96X-at- {')E {"!O;FQY
M!P {'!P {'("`-at-0V]M {&QE=&4-at-4T5-+"!N;VXM8FEO;` {'!P {'!R`-at-("`-at-("!3
M {&5C:6UE;B!P {F5P(&]N;'D'!P {'!P=.96=A=&EV92!S=&%I;B`H8V]M {&QE
M=&4I!P {'!P {'3F5G871I=F4-at- {W1A:6X-at-*'!R97`-at-;VYL} 2D'!P {'!P=& {F5E
M} F4-at-9G)A8W1U {F4-at-*'!R97`-at-;VYL} 2D'!P {'!P=& {F5E} F4-at-9G)A8W1U {F4-at-
M:6YS='(-at-=7-E+VAR!P {'!P {'14T-at-=&5C:&YI8V%L(&AE;'`O:'(',3$P+C`P
M!P {'!P {-*FEN8VQU9&5S(&]P97)A=&]R#0U#;VUP:6QE9"!B} 2!$ {BX-at-07)T
M:'5R(%(N($AA;F0L($-E;G1R86P-at-14T-at-1F%C:6QI='DL(%5N:79E {G-I='D-at-
M;V8-at-0V]N;F5C=&EC=70-at-2&5A;'1H($-E;G1E {-at-T&`*;0`J?0`GP```-at-``P``
M`$-at-`2``````"V-at-(H_^'_X0+Y`D4#1P4H`_P``-at-```$-at-`2``````"V`(H``$`
M``!D`````0`#`P,````!)P\``0`!````````````````8`-at-`&0&0````````
M`````````````0````````````````````#0SQ'-at-H;$:X0``````````````
M````````.P`#__X`"0`&``````````````````$````!````````$``````"
M`````?____X``````````/______________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M________________________________________!0!3`'4`;0!M`&$` {-at-!Y
M`$D`;-at-!F`&\` {-at-!M`&$`=`!I`&\`;-at-``````````````````````````````
M`````"-at-``-at-#_______________\`````````````````````````````````
M```````````````"````W-at-$`````````````````````````````````````
M````````````````````````````````````````````````````````````
M`/_______________P``````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````________
M________````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````````#_______________\`
M````````````````````````````````````````````````````````````
M````````0IJ"BQR[`0,``````````P````````!````````````````} ````
M%0```$UI8W)O {V]F="!7;W)D(#8N,"XQ``,`````````'-at-````(````R``,`
M``````````/__-at-`)``8``````````````````0````$````````0``````'_
M___^``````````#_____________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M_________________________P`#```^`P``/P,``*,#``!5!-at-``NP8``/`(
M``!2"0``T0L``#,,``"^#-at-``'-at-\``.,1``!G$-at-``_-at-#^`/X`_-at-#^`/X`_```
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````)U`0`"58$-``,```$#```^
M`P``/P,``$`#``!;`P`` {P,``'T#``"%`P``H-at-,``*,#``"X`P``N0,``+H#
M``"[`P``O`,``+T#``#+`P``T0,``- {#``#=`P``Z0,``.H#``#V`P``_`,`
M``,$```$!```!00``/X``3`J\`#^``$P*O``_-at-`!,"KP`/H``0\+\`#Z``)+
M!_``^-at-`"A-at-;P`/H`` {P$\`#Z``',!/``^-at-`"/-at-?P`,L``CX'\`#(``$/"_``
MR``!2P?P`,-at-``88&\`#(``',!/``R``!/-at-?P`)\``3X'\`#(``$/"_``FP`!
M2P?P`)L``88&\`";``',!/``FP`!/-at-?P`)\``3X'\`#(``$/"_``FP`!2P?P
M`)L``88&\`";``',!/``FP`!/-at-?P``````````````,```4"&`$`*```&`$9
M`;AL`,"C`!,`OD``!0``YPL*%&-at-;#"$B*0``20!*`$D`20```$D`20!)`$D`
M``!)`$D`20!)````20!)`$D`20```$D`20!)`$H```(``!-at-!`"X``!-at-!&0&X
M;`#`HP`3`+Y```4``. {+"A1H&PPA(BD``$H`2-at-!)`$D```!*`$D`20!)````
M2-at-!)`$D`20```$H`20!)`$D```!*`$D`20!*`+\*$!D0&1`9$!D0&0`#```%
M`1-at-!``$``!L%!```!-at-0``"$$```G!```*`0``"D$```N!```+P0``$($``!#
M!```1`0``$4$``!&!```1P0``% {$``!} !```900``&L$``!L!```;00``(L$
M``"1!```EP0``)T$``"} !```GP0``+8$``"\!```P-at-0``,-at-$``#)!```R-at-0`
M`-\$``#S!```^00``/H$``#[!```_`0``- {``3X'\`#4``$/"_``T``!2P?P
M`-```88&\`#0``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-``
M`88&\`#0``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&
M\`#0``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&\`#0
M``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&\`#0``',
M!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&\`#0``',!/``
MT``!/-at-?P`- {``3X'\`````````,```4"&`$``-at-``&`$`*```&`$9`;AL`,"C
M`!,`OD``!0``YPL*%&-at-;#"$B*0``20!*`$D`20```$D`20!)`$D```!)`$D`
M20!)````20!)`$D`20```$D`20!)`$H`)?P$```8!0``'P4``"`%```F!0``
M)P4``"-at-%``!"!0``2`4``$X%``!/!0``4`4``%$%``!K!0`` {04``'(%``!S
M!0``=`4``'4%``"/!0``E04``)L%``" {!0``G04``)X%``"X!0``O-at-4``+\%
M``#%!0``Q-at-4``, {%``#B!0``Z`4``.D%``#J!0``ZP4``.P%```(!-at-``_0`!
M#POP`/D``4L'\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP
M`/D``4L'\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D`
M`4L'\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D``4L'
M\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D``4L'\`#Y
M``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D``4L'\`#Y``&&
M!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP````````*```&`$9`;AL
M`,"C`!,`OD``!0``YPL*%&-at-;#"$B*0``20!*`$D`20```$D`20!)`$D```!)
M`$D`20!)````20!)`$D`20```$D`20!)`$H```,```4"&`$``-at-``&`$E"`8`
M``\&```0!-at-``$08``!(&```3!-at-``,`8``#$&```R!-at-``,P8``#0&```U!-at-``
M2-at-8``$L&``!,!-at-``308``%,&``!4!-at-``508``%8&``!T!-at-``BP8``/P``4L'
M\`#\``&&!O``_``!S`3P`/P``3X'\`#3``$^!_``T``!#POP`/P``4L'\`#\
M``&&!O``_``!S`3P`/P``3X'\`"G``$^!_``T``!#POP`/P``4L'\`#\``&&
M!O``_``!S`3P`/P``3X'\`!^``$^!_``?``!,"KP`-```0\+\`!X``)*!_``
M} ``"I`7P`````````P``!0$8`0`!````*```&`$9`;AL`,"C`!,`OD``!0``
MYPL*%&-at-;#"$B*0``20!*`$H`20```$D`20!*`$D```!)`$D`2-at-!)````20!)
M`$H`20```$D`20!*`$H``"-at-``!-at-!&0&X;`#`HP`3`+Y```4``. {+"A1H&PPA
M(BD``$D`2-at-```$D```!)`$D```!)````20!)````20```$D`20```$D```!)
M`$D```!*```"```8`0`H```8`1D!N&P`P*,`$P"^0``%``#G"PH4:!L,(2(I
M``!)`$H`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!)````20!)
M`$D`2-at-```P``!0(8`16+!-at-``J`8``+0&``"Z!-at-``NP8``-`&``#1!-at-``T-at-8`
M`-,&``#4!-at-``U08``.,&``#O!-at-``]08``/L&```!!P```-at- {```X'```;!P``
M'` {``!T'```} !P``'P {``#H'```[!P``/` {``#T'```^!P``/P {``/P``M`&
M\`#\``$=!O``_``!'0;P`,T``1T&\`#*``$/"_``R-at-`!2-at-?P`,H``:0%\`#*
M``'0!O``R-at-`!'0;P`*$``1T&\`#*``$/"_``G0`!2-at-?P`)T``:0%\`"=``'0
M!O``G0`!'0;P`*$``1T&\`#*``$/"_``G0`!2-at-?P`)T``:0%\`"=``'0!O``
MG0`!'0;P`*$``1T&\`#*``$/"_``G0`!2-at-?P`)T``:0%\`"=``'0!O``G0`!
M'0;P`*$``1T&\`````,```4"&`$`*```&`$9`;AL`,"C`!,`OD``!0``YPL)
M%(4:+2(B*0``20!*`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`
M20```$D`20!)`$H```(``!-at-!`"X``!-at-!&0&X;`#`HP`3`+Y```4``. {+"12%
M&BTB(BD``$H`2-at-!)`$D```!*`$D`20!)````2-at-!)`$D`20```$H`20!)`$D`
M``!*`$D`20!*`+\*$!D0&1`9$!D0&0`#```%`1-at-!'#\'``!2!P``4P {``%0'
M``!5!P``5-at- {``% {'``!G!P``:` {``&D'``!J!P``:P {``&P'``"*!P``BP {`
M`(P'``"-!P``C-at- {``(\'``"F!P``IP {``*-at-'``"I!P``J-at- {``*L'``#`!P``
MP0 {``,('``##!P``Q` {``,4'``#A!P``X-at- {``.,'``#D!P``Y0 {``.8'````
M"```_0`!#POP`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``
M_0`!#POP`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!
M#POP`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP
M`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP`/D`
M`4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP`/D``4H'
M\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP````````*```
M&`$9`;AL`,"C`!,`OD``!0``YPL)%(4:+2(B*0``20!*`$D`20```$D`20!)
M`$D```!)`$D`20!)````20!)`$D`20```$D`20!)`$H```,```4"&`$``-at-``
M&`$E``-at-```$(```""````P-at-```0(```%"```'P-at-``"`(```A"```(-at--at-``",(
M```D"```/-at--at-``#\(``!`"```00-at-``$((``!#"```70-at-``%X(``!?"```8`-at-`
M`&$(``!B"```?0-at-``'X(``!_"```-at-`-at-``($(``"""```G-at--at-``)\(``"-at-"```
MH0-at-``*((``"C"```P`-at-``,$(``#\``%*!_``_``!I`7P`/P``=`&\`#\``$=
M!O``TP`!'0;P`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``
MTP`!'0;P`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!
M'0;P`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!'0;P
M`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!'0;P`-``
M`0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!'0;P`-```0\+
M\`#\``%*!_`````````"```8`0`H```8`1D!N&P`P*,`$P"^0``%``#G"PD4
MA1HM(B(I``!)`$H`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!)
M````20!)`$D`2-at-```P``!0(8`27!"```P-at--at-``,,(``#$"```Q0-at-``-H(``#-at-
M"```X0-at-``. {(``#M"```[-at--at-``.\(``#P"```\0-at-```H)```A"0``.`D``%$)
M``!2"0``9PD``/P``:0%\`#\``'0!O``_``!'0;P`-,``1T&\`#0``$/"_``
M_``!2-at-?P`/P``:0%\`#\``'0!O``_``!'0;P`* {``1T&\`"E``$P*O``H-at-`!
M,"KP`-```7P+\`"} ``*!!O``G-at-`"FP?P`)X``GP%\`"} ``**!O``;P`"B-at-;P
M`-```7P+\``````````````N```8`1D!N&P`P*,`$P"^0``%``!4#*T3(!QT
M(M8I``!*`$H`20!)````2-at-!)`$D`20```$H`20!)`$D```!*`$D`20!)````
M2-at-!)`$D`2-at-"_"A`9$!D0&1`9$!D``P``!0$8`0`"```)`0`!````*```&`$9
M`;AL`,"C`!,`OD``!0``YPL)%(4:+2(B*0``20!*`$H`20```$D`20!*`$D`
M``!)`$D`2-at-!)````20!)`$H`20```$D`20!*`$H```(``!-at-!`"-at-``!-at-!&0&X
M;`#`HP`3`+Y```4``. {+"12%&BTB(BD``$D`2-at-```$D```!)`$D```!)````
M20!)````20```$D`20```$D```!)`$D```!*```#```%`A-at-!$V {)``!H"0``
M:0D``&H)``!K"0``;`D``'H)``"`"0``A-at-D``(P)``"8"0``F0D``*4)``"F
M"0``IPD``*-at-)``"I"0``J-at-D``,4)``#+"0``S`D``-()``#3"0``U`D``. {)
M``#H"0``Z0D``.H)``#K"0``[`D``/P)```#"-at-``!`H```H*```+"-at-``#`H`
M`"H*```P"-at-``_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0
M``%\"_``_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\
M"_``_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``
M_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``_``!
M-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``_``!-at-0;P
M`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``_``!-at-0;P````
M`````-at-``&`$`*```&`$9`;AL`,"C`!,`OD``!0``5`RM$R` {="+6*0``20!*
M`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`20```$D`20!)`$H`
M``,```4"&`$E,`H``#8*```\"-at-``/0H``#X*``!5"-at-``6PH``&$*``!G"-at-``
M:`H``&D*``!^"-at-``?PH``(`*``"&"-at-``APH``(-at-*``"D"-at-``JPH``*P*``"M
M"-at-``K-at-H``*\*``#)"-at-``SPH``-4*``#;"-at-``W`H``-T*``#W"-at-``^`H``/D*
M``#Z"-at-``^PH``/P*```6"P``%PL``!-at-+``#\``&;!_``_``!?`7P`/P``8H&
M\`#3``&*!O``T``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3
M``&*!O``T``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*
M!O``T``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``
MT``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!
M?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP
M`/P``8$&\`#\``&;!_`````````"```8`0`H```8`1D!N&P`P*,`$P"^0``%
M``!4#*T3(!QT(M8I``!)`$H`20!)````20!)`$D`20```$D`20!)`$D```!)
M`$D`20!)````20!)`$D`2-at-```P``!0(8`248"P``'-at-L``!\+```-at-"P``.-at-L`
M`#L+```\"P``/0L``#X+```_"P``6-at-L``%L+``!A"P``8-at-L``&,+``!D"P``
M-at-`L``($+``"'"P``B`L``(D+``"*"P``IPL``*-at-+``"N"P``KPL``+`+``"Q
M"P``Q-at-L``, {+``#("P``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P`
M`8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P``8$&
M\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P``8$&\`#\
M``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P``8$&\`#\``&;
M!_``_``!?`7P`/P``8H&\`"G``&*!O``T``!?`OP`/P``8$&\`#\``&;!_``
M*```&`$9`;AL`,"C`!,`OD``!0``5`RM$R` {="+6*0``20!*````20```$D`
M20```$D```!)`$D```!)````20!)````20```$D`20```$H```(``!-at-!`"-at-`
M`!-at-!&0&X;`#`HP`3`+Y```4``%0,K1,-at-''0BUBD``$D`2-at-!)`$D```!)`$D`
M20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!*```#```%`A-at-!'L-at-+
M``#)"P``SPL``-`+``#1"P``T-at-L``.P+``#X"P```-at-P```X,```5#```)`P`
M`#(,```S#```2`P``$D,``!*#```2PP``$P,``!-#```6PP``/P``7P%\`#\
M``&*!O``TP`!B-at-;P`-$``3`J\`#.``$0"_``R-at-`"*-at-?P`,H``5`&\`#*``%0
M!O``R-at-`!B-at-;P`,H``8H&\`#*``&*!O``R-at-`!B-at-;P`)L``8H&\`#.``$0"_``
M_``!*-at-?P`/P``5`&\`#\``&*!O``_``!B-at-;P`'(``8H&\`#.``$0"_``````
M`"-at-``!-at-!&0&X;`#`HP`3`+Y```4``.-at-+ZA,2&W0BUBD``$D`2-at-!)`$D```!)
M`$D`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!*```N```8`1D!
MN&P`P*,`$P"^0``%``#H"^H3$AMT(M8I``!*`$H`20!)````2-at-!)`$D`20``
M`$H`20!)`$D```!*`$D`20!)````2-at-!)`$D`2-at-"_"A`9$!D0&1`9$!D``P``
M!0$8`0`"```8`0`!````*```&`$9`;AL`,"C`!,`OD``!0``5`RM$R` {="+6
M*0``20!*`$H`20```$D`20!*`$D```!)`$D`2-at-!)````20!)`$H`20```$D`
M20!*`$H```,```4"&`$46PP``&$,``!G#``` {PP``'D,``!Z#```A-at-P``( {,
M``"(#```B0P``)`,``"1#```K`P``+$,``"W#```R0P``,H,``#+#```W-at-P`
M`-\,``#-at-#```X0P``.(,``#C#```\PP``/0,``#[#```_`P``/T,``#^#```
M'`T``"(-```C#0``)`T``"4-```F#0``/0T``#X-``#\``$J!_``_``!4`;P
M`/P``8H&\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P`
M`8H&\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`-```8H&
M\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\
M``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*
M!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*!O``
MTP`!B-at-;P`-```1`+\`#\``$J!_`````````"```8`0`H```8`1D!N&P`P*,`
M$P"^0``%``#H"^H3$AMT(M8I``!)`$H`20!)````20!)`$D`20```$D`20!)
M`$D```!)`$D`20!)````20!)`$D`2-at-```P``!0(8`24^#0``1`T``$4-``!&
M#0``1PT``%P-``!=#0``7-at-T``%\-``!-at-#0``80T``'T-``!^#0``A`T``(4-
M``"�``APT``*$-``"R#0``LPT``+0-``"U#0``M-at-T``-`-``#1#0``T-at-T`
M`-,-``#4#0``U0T``.\-``#P#0``\0T``/(-``#S#0``]`T```X.```/#-at-``
M$`X``/P``5`&\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P
M`/P``5`&\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P`
M`5`&\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``T``!*-at-?P`/P``5`&
M\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P``5`&\`#\
M``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P``5`&\`#\``&*
M!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P``5`&\`````````(`
M`!-at-!`"-at-``!-at-!&0&X;`#`HP`3`+Y```4``.-at-+ZA,2&W0BUBD``$D`2-at-!)`$D`
M``!)`$D`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!*```#```%
M`A-at-!)1`.```1#-at-``$-at-X``!,.```N#-at-``,PX``#0.```U#-at-``--at-X``# {.``!3
M#-at-``5`X``%4.``!6#-at-``5PX``%-at-.``!U#-at-``=-at-X``' {.``"*#-at-``BPX``(P.
M``"A#-at-``IPX``+D.``"Z#-at-``NPX``/P``8H&\`#\``&*!O``TP`!B-at-;P`-``
M`1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*!O``TP`!B-at-;P`-```1`+
M\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\
M``$J!_``_``!4`;P`-```8H&\`#\``&*!O``IP`!B-at-;P`-```1`+\`#\``$J
M!_``T``!4`;P`/P``8H&\`#\``&*!O``````````````````````````````
M```````````````````````````````H```8`1D!N&P`P*,`$P"^0``%``#H
M"^H3$AMT(M8I``!)`$H```!)````20!)````20```$D`20```$D```!)`$D`
M``!)````20!)````2-at-```-at-``&`$`*```&`$9`;AL`,"C`!,`OD``!0``Z`OJ
M$Q(;="+6*0``20!*`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`
M20```$D`20!)`$H```,```4"&`$:NPX``+P.``"]#-at-``O-at-X``+\.``#)#-at-``
MV0X``.$.``#N#-at-``^0X```$/```.#P``'0\``!X/```S#P``-`\``#4/```V
M#P``-P\``- {``8H&\`#5``$P*O``T-at-`!,"KP`,\``;8*\`#*``&D!_``R-at-`!
MI`?P`,H``3`&\`#*``$P!O``R-at-`!(-at-7P`,H``2(%\`#*``&8!_``R-at-`!F`?P
M`)L``9-at-'\`#/``&V"O``E-at-`!I`?P`)(``3`&\`"6``$B!?``E-at-`!F`?P````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````````````````````````,`
M``4"&`$`!```!0()`1-at-!`"X``!-at-!&0&X;`#`HP`3`+Y```4``(X+"A02&PPA
M?"D``$H`2-at-!)`$D```!*`$D`20!)````2-at-!)`$D`20```$H`20!)`$D```!*
M`$D`20!*`+\*$!D0&1`9$!D0&0`$```%`0D!&`$``-at-``&`$``-at-``"0$``0``
M`"-at-``!-at-!&0&X;`#`HP`3`+Y```4``.-at-+ZA,2&W0BUBD``$D`2-at-!*`$D```!)
M`$D`2-at-!)````20!)`$H`20```$D`20!*`$D```!)`$D`2-at-!*`!(W#P``.`\`
M`$8/``!,#P``6`\``&0/``!E#P``9-at-\``'(/``"!#P``-at--at-\``(,/``"$#P``
MA0\``*`/``"A#P``H-at-\``*,/``"D#P``I0\``+-at-/``"Y#P``N-at-\``+L/``"\
M#P``O0\``,T/``#.#P``SP\``-`/``#E#P``Y-at-\```00```%$```!A````L0
M```,$```UP`!F`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8
M!_``UP`!F`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_``
MUP`!F`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_``UP`!
MF`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_``UP`!F`?P
M`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#+``&8!_``UP`!F`?P`-0`
M`;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_```````````P``"0$8
M`0`$```%`-at-D!&`$``-at-``&`$`*```&`$9`;AL`,"C`!,`OD``!0``C-at-L*%!(;
M#"%\*0``20!*`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`20``
M`$D`20!)`$H`)`P0```-$```)!```"40```F$```-1```#80```W$```3!``
M`$T0``!.$```3Q```%`0``!1$```;1```&X0``!O$``` {!```'$0``!R$```
MC!```(T0``".$```CQ```)`0``"1$```JQ```*P0``"M$```KA```*\0``"P
M$```RA```,L0``#,$```S1```- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&
M\`#+``$B!?``SP`!F`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/
M``$B!?``SP`!F`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/``$B
M!?``SP`!F`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/``$B!?``
MSP`!F`?P`- {``9-at-'\`#4``&V"O``QP`!I`?P`,\``3`&\`#/``$B!?``SP`!
MF`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/``$B!?``````````
M`````P``!0(8`0`#```)`1-at-!``0```4""0$8`0`"```8`0`H```8`1D!N&P`
MP*,`$P"^0``%``"."PH4$AL,(7PI``!)`$H`20!)````20!)`$D`20```$D`
M20!)`$D```!)`$D`20!)````20!)`$D`2-at-`CS1```,X0``#/$```Z1```.H0
M``#K$```[!```.T0``#N$```"1$```H1```+$0``#!$```T1```.$0``*A$`
M`"L1```L$0``+1$``"X1```O$0``3!$``$T1``!.$0``3Q$``%`1``!1$0``
M9A$``&T1``!N$0``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&
M\`#[``$B!?``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\`#[
M``$B!?``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\`#[``$B
M!?``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\`#[``$B!?``
M^P`!F`?P`*8``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\``````````````H
M```8`1D!N&P`P*,`$P"^0``%``"."PH4$AL,(7PI``!)`$H```!)````20!)
M````20```$D`20```$D```!)`$D```!)````20!)````2-at-```-at-``&`$`*```
M&`$9`;AL`,"C`!,`OD``!0``C-at-L*%!(;#"%\*0``20!*`$D`20```$D`20!)
M`$D```!)`$D`20!)````20!)`$D`20```$D`20!)`$H```0```4""0$8`1UN
M$0``;Q$``'`1``!Q$0`` {A$``(41``"&$0``XQ$``/L``2(%\`#[``&8!_``
MT-at-`!F`?P`-```3`J\`#0``$P*O``T``!,"KP`-```3`J\```````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M``````````````$````H```8`1D!N&P`P*,`$P"^0``%``"."PH4$AL,(7PI
M``!)`$H`2-at-!)````20!)`$H`20```$D`20!*`$D```!)`$D`2-at-!)````20!)
M`$H`2-at-``!```!0()`1-at-!!PX`#P`(``$`2P`/`````P` {``!`\?\"`!P`!DYO
M {FUA;``"````!-at-!A"01C&``````````````````````````B`$%`\O^A`"(`
M%D1E9F%U;'0-at-4&%R86=R87!H($9O;G0```````````````````#C#-at-``!`#C
M$0``!`#K$0```P`$(?__`0``(?__`-at-`$(/__`P``````[P4``+T+``#C#-at-``
M```!`````0`!`````-at-````````,``& {2```*```#```%!```_`0```-at-&``"+
M!-at-``/P {````(``#!"```9PD``#`*```8"P``R`L``%L,```^#0``$`X``+L.
M```W#P``#!```,T0``!N$0``XQ$```L`#``-``X`#P`0`!$`$-at-`3`!0`%0`6
M`! {`&``9`!H`&P` {`!T`'-at-"I`B1! {G1H=7(-at-4BX-at-2&%N9"!A;F0-at-36%I:F$-at-
M22X-at-365D;FEE:W,505)((#DV.D9E92!#;VUP87)I {V]N#D%R=&AU {B!2+B!(
M86YD+DUA8VEN=&]S:"!(1#I496UP;W)A {GD-at-271E;7,Z5V]R9"!7;W)K($9I
M;&4-at-02`.07)T:'5R(%(N($AA;F0505)((#DV.D9E92!#;VUP87)I {V]N#D%R
M=&AU {B!2+B!(86YD0DUA8VEN=&]S:"!(1#I-:6-R;W-O9G0-at-3V9F:6-E.DUI
M8W)O {V]F="!7;W)D(#8Z0T5-1CI&964-at-0V]M {&%R:7-O;-at-Y! {G1H=7(-at-4BX-at-
M2&%N9"]-86-I;G1O {V-at--at-2$0Z5&5M {&]R87)Y($ET96US.E=O {F0-at-5V]R:R!&
M:6QE($$-at-,PY! {G1H=7(-at-4BX-at-2&%N9"]-86-I;G1O {V-at--at-2$0Z5&5M {&]R87)Y
M($ET96US.E=O {F0-at-5V]R:R!&:6QE($$-at-,PY! {G1H=7(-at-4BX-at-2&%N9"]-86-I
M;G1O {V-at--at-2$0Z5&5M {&]R87)Y($ET96US.E=O {F0-at-5V]R:R!&:6QE($$-at-,PY!
M {G1H=7(-at-4BX-at-2&%N9$)-86-I;G1O {V-at--at-2$0Z36EC {F]S;V9T($]F9FEC93I-
M:6-R;W-O9G0-at-5V]R9"`V.D-%348Z1F5E($-O;7!A {FES;VX.07)T:'5R(%(N
M($AA;F1"36%C:6YT;W-H($A$.DUI8W)O {V]F="!/9F9I8V4Z36EC {F]S;V9T
M(%=O {F0-at--CI#14U&.D9E92!#;VUP87)I {V]N#D%R=&AU {B!2+B!(86YD1DUA
M8VEN=&]S:"!(1#I-:6-R;W-O9G0-at-3V9F:6-E.DUI8W)O {V]F="!7;W)D(#8Z
M0T5-1CI&964-at-0V]M {&%R:7-O;BU-86/_0`.``0````````````-at-``0``````
M```"``````````$``0"O`M8!/0`5!-at-``30!4:6UE {R!.97 {-at-4F]M86X`#`8`
M``(`4WEM8F]L``L&``!-`$%R:6%L``L&D`%-`%1I;65S`"(`!``Q"(P8``#0
M`-at-``:`$`````;^(#AF_B`X8``````-at-`````````````````````````$`(,0
M```````````````````````````````````A`P````!*````&D-/35!!4DE3
M3TX-at-3T8-at-14T-at-55-%4B!&1453````#D%R=&AU {B!2+B!(86YD#D%R=&AU {B!2
M+B!(86YD`````````````-#/$} "AL1KA```````````````````````[``/_
M_-at-`)``8``````````````````0````$````````0``````(````!_____-at-``
M````````____________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
6________________________________
`
end

--=====================_19263778==_
Content-Type: application/octet-stream; name="Fee Comparison-PC"
Content-Transfer-Encoding: x-uuencode

begin 666 Fee Comparison-PC
MT,\1X*&Q&N$`````````````````````.P`#`/[_"0`&```````````````!
M`````0``````````$````-at-````$```#^____``````````#_____________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M_______________________]____#-at-```/[___\/````!0````8````'````
M"`````D````*````"P````P````-````$````/[____^____$0```!(````3
M````%````!4````6````%P```!-at-````9````&-at-```!L```` {````'0```!X`
M```?````(````"$````B````(P```"0````E````)-at-```" {```#^________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M_____________________________________________P!2````````````
M````````````````````````````````````````````````````````````
M```````````"``4`__________\#``````D"``````#`````````1-at-``````
M``````````#E} Y2+'+L!`P```(`"`````````0!#`&\`;0!P`$\`8-at-!J````
M````````````````````````````````````````````````````````````
M`!(``-at-'_______________\`````````````````````````````````````
M````````````````7`````````!7`&\` {-at-!D`$0`;P!C`'4`;0!E`&X`=```
M````````````````````````````````````````````````````&-at-`"`?__
M__\$````_____P``````````````````````````````````````````````
M``0````&0P```````$\`8-at-!J`&4`8P!T`%``;P!O`&P`````````````````
M```````````````````````````````````````````6``$!`0````(```#_
M____````````````````````````````Y7N4BQR[`0#E} Y2+'+L!````````
M`````````0```/[___\#````!`````4````&````!P````-at-````)````_O__
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M______________________________\!`/[_`-at-`!`/____\`"0(``````,``
M``````!&'````$UI8W)O {V]F="!7;W)D(#8N,"!$;V-U;65N=`#^____3D(V
M5Q````!7;W)D+D1O8W5M96YT+C8````````[``/__-at-`)``8`````````````
M`````0````$````````0`/[_```#"-at-$```````````````````````$```#R
MGX7-at-3_D0:*N1"``K)[/9,````*X!```/````!P```)-at-````"````X`````0`
M```#`0``"````!H!```!````,0$```P````W`0``"P```$,!```-````3P$`
M``\```!;`0``$````&,!```*````:P$``!(```!W`0``#-at-```)0!```)````
MG`$``!,```"F`0``________________________________'-at-```$````!-
M86-I;G1O {V-at--at-2$0Z36EC {F]S;V9T($]F9FEC93I-:6-R;W-O9G0-at-5V]R9"`V
M.E1E;7!L871E {SI.;W)M86P`'-at-```!L```!#3TU005))4T].($]&($5-(%53
M15(-at-1D5%4P`} ````#P```$%R=&AU {B!2+B!(86YD`!X````/````07)T:'5R
M(%(N($AA;F0``-at-```!`G0`````!"FH*+'+L!0`````"`=M/RS. {!0-RE:`!/
MP`D$`````&4```````$```$``0`#``#C$0``!D,`````````````````````
M``#C#-at-``````````````````````````````````````````````/-at-``;```
M```^``!L````;#X```````!L/-at-```````&P^````````;#X```````!L/-at-``
M%````+8^````````MCX```````"V/-at-```````+8^````````MCX```````"V
M/-at-``"-at-```,`^``!\````MCX````````G0-at-``/0```#P_````````/#\`````
M```\/P```````#P_````````/#\````````\/P```````#P_````````/#\`
M``````#_00```-at-````%"`````````4(````````!0-at-````````%"````````
M`4(````````!0-at-``)-at-```&1"``!8````O$(``$H````G0-at-``````````````
M``````````````!L/-at-```````#P_```````````*``L``0`4`#P_````````
M/#\`````````````````````````````/#\````````\/P```````"="````
M````/#\```````!L/-at-```````&P^````````/#\`````````````````````
M````````/#\````````\/P```````#P_````````/#\````````\/P``````
M`&P^````````/#\```````!L/-at-```````#P_````````_T$`````````````
M````````````````-at-#X``!0```"4/-at-``(-at-```&P^````````;#X```````!L
M/-at-```````&P^````````/#\```````#_00```````#P_``##`-at-``/#\`````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````````````````````````U#
M3TU005))4T].($]&($5-(%5315(-at-1D5%4R!&3U(-at-4T5,14-4140-at-4T525DE#
M15,L($UA {F-H(#$Y.38-#0=#96YT {F%L($5-($9A8VEL:71Y+"!50V]N;-at-=5
M($UI8V-at--at-4V-H(&]F($1E;G1I {W1R} 0=5($9L;W)I9&$-14T-at-0V]R90=5($AA
M=V%I:2P-at-4&%C($)I;VUE9"!297,-at-0W1R!P=414TL(%-%32!U {V4L('!E {B!H
M {-at- {'!P {'!R`-at-(%5N:79E {G-I='D'-3`N,#`',S`N,#`',C4N,#`',S(N,#`M
M,S8N,#`'!R`-at-($EN9'5S=')Y!S4P+C`P!S$P,"XP,` {'!P=5;'1R86UI8W)O
M=&]M92!U {V4L('!E {B!H {-at- {Q,"XP,` {'!S8N,#`'!U-A;7!L92!P {F5P87)A
M=&EO;-at- {'!P {'!R`-at-($-O;7!L971E(%1%30 {Q,3`N,#`',34P+C`P!S {U+C`P
M!P {'("`-at-("`-at-($9I} &%T:6]N(&%N9"!E;6)E9&1I;F {',S`N,#`',3`N,#`'
M,C4N,#`'!P {-at-("`-at-("`-at-5&AI;B!S96-T:6]N:6YG!S4P+C`P!S$R+C4P!S(U
M+C`P!P {'("`-at-("`-at-($=R:60-at- {W1A:6YI;F {'*&EN8VQU9&5D(&EN(&%B;W9E
M*0 {Q,"XP,` {'!P {-at-("!#;VUP;&5T92!314TL(&)I;VQO9VEC86P',3(P+C`P
M!P {W-2XP,` {'!R`-at-("`-at-("!3 {&5C:6UE;B!P {F5P(&]N;'D'-3`N,#`'-#`N
M,#`'!P {'("`-at-0V]M {&QE=&4-at-4T5-+"!N;VXM8FEO;` {X,"XP,` {'!P {'("`-at-
M("`-at-(%-P96-I;65N('!R97`-at-;VYL} 0 {S,"XP,` {R,"XP,` {'!P=.96=A=&EV
M92!S=&%I;B`H8V]M {&QE=&4I!S4P+C`P!P {T,"XP,` {'!TYE9V%T:79E('-T
M86EN("AP {F5P(&]N;'DI!S(P+C`P!P {'!P=& {F5E} F4-at-9G)A8W1U {F4-at-*'!R
M97`-at-;VYL} 2D',3`P+C`P!P {'!P=& {F5E} F4-at-9G)A8W1U {F4-at-:6YS='(-at-=7-E
M+VAR!P {'!P {'14T-at-=&5C:&YI8V%L(&AE;'`O:'('!P {',C4N,#`'!PT'3VAI
M;R!3=&%T92!5($5L96-T($]P=&EC {R!&86,'2DA5($5A {G1H("8-at-4&QA;F5T
M(%-C:0=2=71G97)S($UA {FEN92`F($-O87-T86P-at-4V-I!U4-at-36ES {V]U {FD-at-
M#5)O;&QA!P=414TL(%-%32!U {V4L('!E {B!H {-at- {'!P {'!R`-at-(%5N:79E {G-I
M='D',34N,#`M,S4N,#`',S4N,#`',S4N,#`'-S4N,#`'!R`-at-($EN9'5S=')Y
M!SDP+C`P+3(P,"XP,` {'!P {'56QT {F%M:6-R;W1O;64-at-=7-E+"!P97(-at-:'('
M!P {'!P=386UP;&4-at- {')E {&%R871I;VX'!P {'!P {-at-("!#;VUP;&5T92!414T'
M!P {'!P {-at-("`-at-("`-at-1FEX871I;VX-at-86YD(&5M8F5D9&EN9P {'!P {'!R`-at-("`-at-
M("!4:&EN('-E8W1I;VYI;F {'!P {'!P {-at-("`-at-("`-at-1W)I9"!S=&%I;FEN9P {'
M!P {'!R`-at-($-O;7!L971E(%-%32P-at-8FEO;&]G:6-A;` {'!P {'!R`-at-("`-at-("!3
M {&5C:6UE;B!P {F5P(&]N;'D'!P {'!P {-at-("!#;VUP;&5T92!314TL(&YO;BUB
M:6]L!P {'!P {'("`-at-("`-at-(%-P96-I;65N('!R97`-at-;VYL} 0 {'!P {'!TYE9V%T
M:79E('-T86EN("AC;VUP;&5T92D'!P {'!P=.96=A=&EV92!S=&%I;B`H {')E
M {"!O;FQY*0 {'!P {'!T9R965Z92!F {F%C='5R92`H {')E {"!O;FQY*0 {'!P {'
M!T9R965Z92!F {F%C='5R92!I;G-T {B!U {V4O:'('!P {'!P=%32!T96-H;FEC
M86P-at-:&5L {"]H {-at- {R-2XP,` {',C`N,#`',C4N,#`'!PT-!U4-at-36%R} 6QA;F0-at-
M0F%L=&EM;W)E($-T} 0=5($UA {W,-at-06UH97)S="!":6]L;V=Y!U-O($EL;&EN
M;VES($-T {B!F;W(-at-14T'0V]R;F5L;"!-870-at-4V-I($9A8VEL:71Y!P=414TL
M(%-%32!U {V4L('!E {B!H {-at- {'!P {'!R`-at-(%5N:79E {G-I='D',S$N,C4',34N
M,#`',3(N-3`',3 {N,#`M,S,N,#`'!R`-at-($EN9'5S=')Y!P {'!P {'56QT {F%M
M:6-R;W1O;64-at-=7-E+"!P97(-at-:'(',3(N-3`'!S$P+C`P!P {'4V%M {&QE('!R
M97!A {F%T:6]N!P {'!P {'("`-at-0V]M {&QE=&4-at-5$5-!S(U,"XP,` {'.34N,#`'
M!P {-at-("`-at-("`-at-1FEX871I;VX-at-86YD(&5M8F5D9&EN9P {T,"XP,` {Q,"XP,` {T
M,"XP,` {'!R`-at-("`-at-("!4:&EN('-E8W1I;VYI;F {',34N,#`',3`N,#`',C`N
M,#`'!P {-at-("`-at-("`-at-1W)I9"!S=&%I;FEN9P {'!S$U+C`P!P {'("`-at-0V]M {&QE
M=&4-at-4T5-+"!B:6]L;V=I8V%L!S(W-2XP,` {'!P {'("`-at-("`-at-(%-P96-I;65N
M('!R97`-at-;VYL} 0 {W-2XP,` {U-2XP,` {T,"XP,` {'!R`-at-($-O;7!L971E(%-%
M32P-at-;F]N+6)I;VP'!P {'!P {-at-("`-at-("`-at-4W!E8VEM96X-at- {')E {"!O;FQY!P {'
M,C`N,#`'!P=.96=A=&EV92!S=&%I;B`H8V]M {&QE=&4I!P {'!P {'3F5G871I
M=F4-at- {W1A:6X-at-*'!R97`-at-;VYL} 2D'!S$P+C`P!P {'!T9R965Z92!F {F%C='5R
M92`H {')E {"!O;FQY*0 {'-#`N,#`'!P {'1G)E97IE(&9R86-T=7)E(&EN {W1R
M('5S92]H {-at- {',C`N,#`'!P {'14T-at-=&5C:&YI8V%L(&AE;'`O:'('!P {',S(N
M,#`'!PT'26]W82!3=&%T92!"97-S97D-at-36EC($9A8P=,;WEO;&$-at-52!-0PU#
M96QL($)I;VP'52!/:VQA:&]M82P-3F]R;6%N!U4-at-0V%L:68L($1A=FES#49A
M8R!!9'8-at-26YS='('!U1%32P-at-4T5-('5S92P-at- {&5R(&AR!P {'!P {'("`-at-56YI
M=F5R {VET} 0 {R-RXU,` {Q,"XP,` {R-"XU,"TS-"XP,` {Q-2XP,` {'("`-at-26YD
M=7-T {GD'!P {',3 {U+C`P!P=5;'1R86UI8W)O=&]M92!U {V4L('!E {B!H {-at- {V
M+C8P!S$P+C`P!RAC {GEO+V1A} 2D-at-(#$U+C`P!P {'4V%M {&QE('!R97!A {F%T
M:6]N!P {'!P {'("`-at-0V]M {&QE=&4-at-5$5-!P {Q,C(N-3`'!P {'("`-at-("`-at-($9I
M} &%T:6]N(&%N9"!E;6)E9&1I;F {',38N-3`'!P {'!R`-at-("`-at-("!4:&EN('-E
M8W1I;VYI;F {'!S$U+C`P!P {'!R`-at-("`-at-("!' {FED('-T86EN:6YG!P {'!P {'
M("`-at-0V]M {&QE=&4-at-4T5-+"!B:6]L;V=I8V%L!P {Y,"XP,` {'!P {-at-("`-at-("`-at-
M4W!E8VEM96X-at- {')E {"!O;FQY!RA&:7-at-F0U!$*2`-at-,38N-3`'!P {'!R`-at-($-O
M;7!L971E(%-%32P-at-;F]N+6)I;VP'!P {'!P {-at-("`-at-("`-at-4W!E8VEM96X-at- {')E
M {"!O;FQY!P {'!P {'3F5G871I=F4-at- {W1A:6X-at-*&-O;7!L971E*0 {'!P {'!TYE
M9V%T:79E('-T86EN("AP {F5P(&]N;'DI!S8N-C`'!P {'!T9R965Z92!F {F%C
M='5R92`H {')E {"!O;FQY*0 {'!P {'!T9R965Z92!F {F%C='5R92!I;G-T {B!U
M {V4O:'('!P {H {&5R(&1A} 2D-at-("`-at-,3`N,#`'!P=%32!T96-H;FEC86P-at-:&5L
M {"]H {-at- {Q-BXU,` {H4V-O {&4I("`-at-("`Q,"XP,` {'!P {-#0=%4$9,+4-)344-
M3&%U {V%N;F4L(%-W:71Z!U1R96YT(%4-3VYT87)I;RP-at-0V%N!U4-at-;V8-at-5'5R
M:W4-1FEN;&%N9`=5;FEV($UE9"!38V-at--1&5B {F5C96XL($AU;F {'!U1%32P-at-
M4T5-('5S92P-at- {&5R(&AR!P {'!P {'("`-at-56YI=F5R {VET} 0 {R,"XP,` {Q,"XP
M,"TQ-2XP,` {Q-RXU,"TT-"XU,` {'!R`-at-($EN9'5S=')Y!RHS,#`N,#`M-3`P
M+C`P!P {'!P=5;'1R86UI8W)O=&]M92!U {V4L('!E {B!H {-at- {'!P {'!U-A;7!L
M92!P {F5P87)A=&EO;-at- {'!P {'!R`-at-($-O;7!L971E(%1%30 {'!P {H=R]O('!H
M;W1O*2`-at-("`-at--RXP,` {'("`-at-("`-at-($9I} &%T:6]N(&%N9"!E;6)E9&1I;F {'
M!P {S+C(U!P {'("`-at-("`-at-(%1H:6X-at- {V5C=&EO;FEN9P {'!RAP97(-at-9W)I9"D-at-
M+C {U!P {'("`-at-("`-at-($=R:60-at- {W1A:6YI;F {'!P {'!P {-at-("!#;VUP;&5T92!3
M14TL(&)I;VQO9VEC86P'!P {'!P {-at-("`-at-("`-at-4W!E8VEM96X-at- {')E {"!O;FQY
M!P {'!P {'("`-at-0V]M {&QE=&4-at-4T5-+"!N;VXM8FEO;` {'!P {'!R`-at-("`-at-("!3
M {&5C:6UE;B!P {F5P(&]N;'D'!P {'!P=.96=A=&EV92!S=&%I;B`H8V]M {&QE
M=&4I!P {'!P {'3F5G871I=F4-at- {W1A:6X-at-*'!R97`-at-;VYL} 2D'!P {'!P=& {F5E
M} F4-at-9G)A8W1U {F4-at-*'!R97`-at-;VYL} 2D'!P {'!P=& {F5E} F4-at-9G)A8W1U {F4-at-
M:6YS='(-at-=7-E+VAR!P {'!P {'14T-at-=&5C:&YI8V%L(&AE;'`O:'(',3$P+C`P
M!P {'!P {-*FEN8VQU9&5S(&]P97)A=&]R#0U#;VUP:6QE9"!B} 2!$ {BX-at-07)T
M:'5R(%(N($AA;F0L($-E;G1R86P-at-14T-at-1F%C:6QI='DL(%5N:79E {G-I='D-at-
M;V8-at-0V]N;F5C=&EC=70-at-2&5A;'1H($-E;G1E {-at-T&`*;0`J?0`GP```-at-``P``
M`$-at-`2``````"V-at-(H_^'_X0+Y`D4#1P4H`_P``-at-```$-at-`2``````"V`(H``$`
M``!D`````0`#`P,````!)P\``0`!````````````````8`-at-`&0&0````````
M`````````````0````````````````````#0SQ'-at-H;$:X0``````````````
M````````.P`#__X`"0`&``````````````````$````!````````$``````"
M`````?____X``````````/______________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M________________________________________!0!3`'4`;0!M`&$` {-at-!Y
M`$D`;-at-!F`&\` {-at-!M`&$`=`!I`&\`;-at-``````````````````````````````
M`````"-at-``-at-#_______________\`````````````````````````````````
M```````````````"````W-at-$`````````````````````````````````````
M````````````````````````````````````````````````````````````
M`/_______________P``````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````________
M________````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````````#_______________\`
M````````````````````````````````````````````````````````````
M````````0IJ"BQR[`0,``````````P````````!````````````````} ````
M%0```$UI8W)O {V]F="!7;W)D(#8N,"XQ``,`````````'-at-````(````R``,`
M``````````/__-at-`)``8``````````````````0````$````````0``````'_
M___^``````````#_____________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
M_________________________P`#```^`P``/P,``*,#``!5!-at-``NP8``/`(
M``!2"0``T0L``#,,``"^#-at-``'-at-\``.,1``!G$-at-``_-at-#^`/X`_-at-#^`/X`_```
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````)U`0`"58$-``,```$#```^
M`P``/P,``$`#``!;`P`` {P,``'T#``"%`P``H-at-,``*,#``"X`P``N0,``+H#
M``"[`P``O`,``+T#``#+`P``T0,``- {#``#=`P``Z0,``.H#``#V`P``_`,`
M``,$```$!```!00``/X``3`J\`#^``$P*O``_-at-`!,"KP`/H``0\+\`#Z``)+
M!_``^-at-`"A-at-;P`/H`` {P$\`#Z``',!/``^-at-`"/-at-?P`,L``CX'\`#(``$/"_``
MR``!2P?P`,-at-``88&\`#(``',!/``R``!/-at-?P`)\``3X'\`#(``$/"_``FP`!
M2P?P`)L``88&\`";``',!/``FP`!/-at-?P`)\``3X'\`#(``$/"_``FP`!2P?P
M`)L``88&\`";``',!/``FP`!/-at-?P``````````````,```4"&`$`*```&`$9
M`;AL`,"C`!,`OD``!0``YPL*%&-at-;#"$B*0``20!*`$D`20```$D`20!)`$D`
M``!)`$D`20!)````20!)`$D`20```$D`20!)`$H```(``!-at-!`"X``!-at-!&0&X
M;`#`HP`3`+Y```4``. {+"A1H&PPA(BD``$H`2-at-!)`$D```!*`$D`20!)````
M2-at-!)`$D`20```$H`20!)`$D```!*`$D`20!*`+\*$!D0&1`9$!D0&0`#```%
M`1-at-!``$``!L%!```!-at-0``"$$```G!```*`0``"D$```N!```+P0``$($``!#
M!```1`0``$4$``!&!```1P0``% {$``!} !```900``&L$``!L!```;00``(L$
M``"1!```EP0``)T$``"} !```GP0``+8$``"\!```P-at-0``,-at-$``#)!```R-at-0`
M`-\$``#S!```^00``/H$``#[!```_`0``- {``3X'\`#4``$/"_``T``!2P?P
M`-```88&\`#0``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-``
M`88&\`#0``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&
M\`#0``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&\`#0
M``',!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&\`#0``',
M!/``T``!/-at-?P`- {``3X'\`#4``$/"_``T``!2P?P`-```88&\`#0``',!/``
MT``!/-at-?P`- {``3X'\`````````,```4"&`$``-at-``&`$`*```&`$9`;AL`,"C
M`!,`OD``!0``YPL*%&-at-;#"$B*0``20!*`$D`20```$D`20!)`$D```!)`$D`
M20!)````20!)`$D`20```$D`20!)`$H`)?P$```8!0``'P4``"`%```F!0``
M)P4``"-at-%``!"!0``2`4``$X%``!/!0``4`4``%$%``!K!0`` {04``'(%``!S
M!0``=`4``'4%``"/!0``E04``)L%``" {!0``G04``)X%``"X!0``O-at-4``+\%
M``#%!0``Q-at-4``, {%``#B!0``Z`4``.D%``#J!0``ZP4``.P%```(!-at-``_0`!
M#POP`/D``4L'\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP
M`/D``4L'\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D`
M`4L'\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D``4L'
M\`#Y``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D``4L'\`#Y
M``&&!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP`/D``4L'\`#Y``&&
M!O``^0`!S`3P`/D``3X'\`#0``$^!_``_0`!#POP````````*```&`$9`;AL
M`,"C`!,`OD``!0``YPL*%&-at-;#"$B*0``20!*`$D`20```$D`20!)`$D```!)
M`$D`20!)````20!)`$D`20```$D`20!)`$H```,```4"&`$``-at-``&`$E"`8`
M``\&```0!-at-``$08``!(&```3!-at-``,`8``#$&```R!-at-``,P8``#0&```U!-at-``
M2-at-8``$L&``!,!-at-``308``%,&``!4!-at-``508``%8&``!T!-at-``BP8``/P``4L'
M\`#\``&&!O``_``!S`3P`/P``3X'\`#3``$^!_``T``!#POP`/P``4L'\`#\
M``&&!O``_``!S`3P`/P``3X'\`"G``$^!_``T``!#POP`/P``4L'\`#\``&&
M!O``_``!S`3P`/P``3X'\`!^``$^!_``?``!,"KP`-```0\+\`!X``)*!_``
M} ``"I`7P`````````P``!0$8`0`!````*```&`$9`;AL`,"C`!,`OD``!0``
MYPL*%&-at-;#"$B*0``20!*`$H`20```$D`20!*`$D```!)`$D`2-at-!)````20!)
M`$H`20```$D`20!*`$H``"-at-``!-at-!&0&X;`#`HP`3`+Y```4``. {+"A1H&PPA
M(BD``$D`2-at-```$D```!)`$D```!)````20!)````20```$D`20```$D```!)
M`$D```!*```"```8`0`H```8`1D!N&P`P*,`$P"^0``%``#G"PH4:!L,(2(I
M``!)`$H`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!)````20!)
M`$D`2-at-```P``!0(8`16+!-at-``J`8``+0&``"Z!-at-``NP8``-`&``#1!-at-``T-at-8`
M`-,&``#4!-at-``U08``.,&``#O!-at-``]08``/L&```!!P```-at- {```X'```;!P``
M'` {``!T'```} !P``'P {``#H'```[!P``/` {``#T'```^!P``/P {``/P``M`&
M\`#\``$=!O``_``!'0;P`,T``1T&\`#*``$/"_``R-at-`!2-at-?P`,H``:0%\`#*
M``'0!O``R-at-`!'0;P`*$``1T&\`#*``$/"_``G0`!2-at-?P`)T``:0%\`"=``'0
M!O``G0`!'0;P`*$``1T&\`#*``$/"_``G0`!2-at-?P`)T``:0%\`"=``'0!O``
MG0`!'0;P`*$``1T&\`#*``$/"_``G0`!2-at-?P`)T``:0%\`"=``'0!O``G0`!
M'0;P`*$``1T&\`````,```4"&`$`*```&`$9`;AL`,"C`!,`OD``!0``YPL)
M%(4:+2(B*0``20!*`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`
M20```$D`20!)`$H```(``!-at-!`"X``!-at-!&0&X;`#`HP`3`+Y```4``. {+"12%
M&BTB(BD``$H`2-at-!)`$D```!*`$D`20!)````2-at-!)`$D`20```$H`20!)`$D`
M``!*`$D`20!*`+\*$!D0&1`9$!D0&0`#```%`1-at-!'#\'``!2!P``4P {``%0'
M``!5!P``5-at- {``% {'``!G!P``:` {``&D'``!J!P``:P {``&P'``"*!P``BP {`
M`(P'``"-!P``C-at- {``(\'``"F!P``IP {``*-at-'``"I!P``J-at- {``*L'``#`!P``
MP0 {``,('``##!P``Q` {``,4'``#A!P``X-at- {``.,'``#D!P``Y0 {``.8'````
M"```_0`!#POP`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``
M_0`!#POP`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!
M#POP`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP
M`/D``4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP`/D`
M`4H'\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP`/D``4H'
M\`#Y``&D!?``^0`!T`;P`/D``1T&\`#0``$=!O``_0`!#POP````````*```
M&`$9`;AL`,"C`!,`OD``!0``YPL)%(4:+2(B*0``20!*`$D`20```$D`20!)
M`$D```!)`$D`20!)````20!)`$D`20```$D`20!)`$H```,```4"&`$``-at-``
M&`$E``-at-```$(```""````P-at-```0(```%"```'P-at-``"`(```A"```(-at--at-``",(
M```D"```/-at--at-``#\(``!`"```00-at-``$((``!#"```70-at-``%X(``!?"```8`-at-`
M`&$(``!B"```?0-at-``'X(``!_"```-at-`-at-``($(``"""```G-at--at-``)\(``"-at-"```
MH0-at-``*((``"C"```P`-at-``,$(``#\``%*!_``_``!I`7P`/P``=`&\`#\``$=
M!O``TP`!'0;P`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``
MTP`!'0;P`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!
M'0;P`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!'0;P
M`-```0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!'0;P`-``
M`0\+\`#\``%*!_``_``!I`7P`/P``=`&\`#\``$=!O``TP`!'0;P`-```0\+
M\`#\``%*!_`````````"```8`0`H```8`1D!N&P`P*,`$P"^0``%``#G"PD4
MA1HM(B(I``!)`$H`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!)
M````20!)`$D`2-at-```P``!0(8`27!"```P-at--at-``,,(``#$"```Q0-at-``-H(``#-at-
M"```X0-at-``. {(``#M"```[-at--at-``.\(``#P"```\0-at-```H)```A"0``.`D``%$)
M``!2"0``9PD``/P``:0%\`#\``'0!O``_``!'0;P`-,``1T&\`#0``$/"_``
M_``!2-at-?P`/P``:0%\`#\``'0!O``_``!'0;P`* {``1T&\`"E``$P*O``H-at-`!
M,"KP`-```7P+\`"} ``*!!O``G-at-`"FP?P`)X``GP%\`"} ``**!O``;P`"B-at-;P
M`-```7P+\``````````````N```8`1D!N&P`P*,`$P"^0``%``!4#*T3(!QT
M(M8I``!*`$H`20!)````2-at-!)`$D`20```$H`20!)`$D```!*`$D`20!)````
M2-at-!)`$D`2-at-"_"A`9$!D0&1`9$!D``P``!0$8`0`"```)`0`!````*```&`$9
M`;AL`,"C`!,`OD``!0``YPL)%(4:+2(B*0``20!*`$H`20```$D`20!*`$D`
M``!)`$D`2-at-!)````20!)`$H`20```$D`20!*`$H```(``!-at-!`"-at-``!-at-!&0&X
M;`#`HP`3`+Y```4``. {+"12%&BTB(BD``$D`2-at-```$D```!)`$D```!)````
M20!)````20```$D`20```$D```!)`$D```!*```#```%`A-at-!$V {)``!H"0``
M:0D``&H)``!K"0``;`D``'H)``"`"0``A-at-D``(P)``"8"0``F0D``*4)``"F
M"0``IPD``*-at-)``"I"0``J-at-D``,4)``#+"0``S`D``-()``#3"0``U`D``. {)
M``#H"0``Z0D``.H)``#K"0``[`D``/P)```#"-at-``!`H```H*```+"-at-``#`H`
M`"H*```P"-at-``_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0
M``%\"_``_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\
M"_``_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``
M_``!-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``_``!
M-at-0;P`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``_``!-at-0;P
M`/P``9L'\`#\``%\!?``_``!B-at-;P`-,``8H&\`#0``%\"_``_``!-at-0;P````
M`````-at-``&`$`*```&`$9`;AL`,"C`!,`OD``!0``5`RM$R` {="+6*0``20!*
M`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`20```$D`20!)`$H`
M``,```4"&`$E,`H``#8*```\"-at-``/0H``#X*``!5"-at-``6PH``&$*``!G"-at-``
M:`H``&D*``!^"-at-``?PH``(`*``"&"-at-``APH``(-at-*``"D"-at-``JPH``*P*``"M
M"-at-``K-at-H``*\*``#)"-at-``SPH``-4*``#;"-at-``W`H``-T*``#W"-at-``^`H``/D*
M``#Z"-at-``^PH``/P*```6"P``%PL``!-at-+``#\``&;!_``_``!?`7P`/P``8H&
M\`#3``&*!O``T``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3
M``&*!O``T``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*
M!O``T``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``
MT``!?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!
M?`OP`/P``8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP
M`/P``8$&\`#\``&;!_`````````"```8`0`H```8`1D!N&P`P*,`$P"^0``%
M``!4#*T3(!QT(M8I``!)`$H`20!)````20!)`$D`20```$D`20!)`$D```!)
M`$D`20!)````20!)`$D`2-at-```P``!0(8`248"P``'-at-L``!\+```-at-"P``.-at-L`
M`#L+```\"P``/0L``#X+```_"P``6-at-L``%L+``!A"P``8-at-L``&,+``!D"P``
M-at-`L``($+``"'"P``B`L``(D+``"*"P``IPL``*-at-+``"N"P``KPL``+`+``"Q
M"P``Q-at-L``, {+``#("P``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P`
M`8$&\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P``8$&
M\`#\``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P``8$&\`#\
M``&;!_``_``!?`7P`/P``8H&\`#3``&*!O``T``!?`OP`/P``8$&\`#\``&;
M!_``_``!?`7P`/P``8H&\`"G``&*!O``T``!?`OP`/P``8$&\`#\``&;!_``
M*```&`$9`;AL`,"C`!,`OD``!0``5`RM$R` {="+6*0``20!*````20```$D`
M20```$D```!)`$D```!)````20!)````20```$D`20```$H```(``!-at-!`"-at-`
M`!-at-!&0&X;`#`HP`3`+Y```4``%0,K1,-at-''0BUBD``$D`2-at-!)`$D```!)`$D`
M20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!*```#```%`A-at-!'L-at-+
M``#)"P``SPL``-`+``#1"P``T-at-L``.P+``#X"P```-at-P```X,```5#```)`P`
M`#(,```S#```2`P``$D,``!*#```2PP``$P,``!-#```6PP``/P``7P%\`#\
M``&*!O``TP`!B-at-;P`-$``3`J\`#.``$0"_``R-at-`"*-at-?P`,H``5`&\`#*``%0
M!O``R-at-`!B-at-;P`,H``8H&\`#*``&*!O``R-at-`!B-at-;P`)L``8H&\`#.``$0"_``
M_``!*-at-?P`/P``5`&\`#\``&*!O``_``!B-at-;P`'(``8H&\`#.``$0"_``````
M`"-at-``!-at-!&0&X;`#`HP`3`+Y```4``.-at-+ZA,2&W0BUBD``$D`2-at-!)`$D```!)
M`$D`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!*```N```8`1D!
MN&P`P*,`$P"^0``%``#H"^H3$AMT(M8I``!*`$H`20!)````2-at-!)`$D`20``
M`$H`20!)`$D```!*`$D`20!)````2-at-!)`$D`2-at-"_"A`9$!D0&1`9$!D``P``
M!0$8`0`"```8`0`!````*```&`$9`;AL`,"C`!,`OD``!0``5`RM$R` {="+6
M*0``20!*`$H`20```$D`20!*`$D```!)`$D`2-at-!)````20!)`$H`20```$D`
M20!*`$H```,```4"&`$46PP``&$,``!G#``` {PP``'D,``!Z#```A-at-P``( {,
M``"(#```B0P``)`,``"1#```K`P``+$,``"W#```R0P``,H,``#+#```W-at-P`
M`-\,``#-at-#```X0P``.(,``#C#```\PP``/0,``#[#```_`P``/T,``#^#```
M'`T``"(-```C#0``)`T``"4-```F#0``/0T``#X-``#\``$J!_``_``!4`;P
M`/P``8H&\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P`
M`8H&\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`-```8H&
M\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\
M``&*!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*
M!O``TP`!B-at-;P`-```1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*!O``
MTP`!B-at-;P`-```1`+\`#\``$J!_`````````"```8`0`H```8`1D!N&P`P*,`
M$P"^0``%``#H"^H3$AMT(M8I``!)`$H`20!)````20!)`$D`20```$D`20!)
M`$D```!)`$D`20!)````20!)`$D`2-at-```P``!0(8`24^#0``1`T``$4-``!&
M#0``1PT``%P-``!=#0``7-at-T``%\-``!-at-#0``80T``'T-``!^#0``A`T``(4-
M``"�``APT``*$-``"R#0``LPT``+0-``"U#0``M-at-T``-`-``#1#0``T-at-T`
M`-,-``#4#0``U0T``.\-``#P#0``\0T``/(-``#S#0``]`T```X.```/#-at-``
M$`X``/P``5`&\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P
M`/P``5`&\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P`
M`5`&\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``T``!*-at-?P`/P``5`&
M\`#\``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P``5`&\`#\
M``&*!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P``5`&\`#\``&*
M!O``_``!B-at-;P`-,``8H&\`#0``$0"_``_``!*-at-?P`/P``5`&\`````````(`
M`!-at-!`"-at-``!-at-!&0&X;`#`HP`3`+Y```4``.-at-+ZA,2&W0BUBD``$D`2-at-!)`$D`
M``!)`$D`20!)````20!)`$D`20```$D`20!)`$D```!)`$D`20!*```#```%
M`A-at-!)1`.```1#-at-``$-at-X``!,.```N#-at-``,PX``#0.```U#-at-``--at-X``# {.``!3
M#-at-``5`X``%4.``!6#-at-``5PX``%-at-.``!U#-at-``=-at-X``' {.``"*#-at-``BPX``(P.
M``"A#-at-``IPX``+D.``"Z#-at-``NPX``/P``8H&\`#\``&*!O``TP`!B-at-;P`-``
M`1`+\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*!O``TP`!B-at-;P`-```1`+
M\`#\``$J!_``_``!4`;P`/P``8H&\`#\``&*!O``TP`!B-at-;P`-```1`+\`#\
M``$J!_``_``!4`;P`-```8H&\`#\``&*!O``IP`!B-at-;P`-```1`+\`#\``$J
M!_``T``!4`;P`/P``8H&\`#\``&*!O``````````````````````````````
M```````````````````````````````H```8`1D!N&P`P*,`$P"^0``%``#H
M"^H3$AMT(M8I``!)`$H```!)````20!)````20```$D`20```$D```!)`$D`
M``!)````20!)````2-at-```-at-``&`$`*```&`$9`;AL`,"C`!,`OD``!0``Z`OJ
M$Q(;="+6*0``20!*`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`
M20```$D`20!)`$H```,```4"&`$:NPX``+P.``"]#-at-``O-at-X``+\.``#)#-at-``
MV0X``.$.``#N#-at-``^0X```$/```.#P``'0\``!X/```S#P``-`\``#4/```V
M#P``-P\``- {``8H&\`#5``$P*O``T-at-`!,"KP`,\``;8*\`#*``&D!_``R-at-`!
MI`?P`,H``3`&\`#*``$P!O``R-at-`!(-at-7P`,H``2(%\`#*``&8!_``R-at-`!F`?P
M`)L``9-at-'\`#/``&V"O``E-at-`!I`?P`)(``3`&\`"6``$B!?``E-at-`!F`?P````
M````````````````````````````````````````````````````````````
M``````````````````````````````````````````````````````````,`
M``4"&`$`!```!0()`1-at-!`"X``!-at-!&0&X;`#`HP`3`+Y```4``(X+"A02&PPA
M?"D``$H`2-at-!)`$D```!*`$D`20!)````2-at-!)`$D`20```$H`20!)`$D```!*
M`$D`20!*`+\*$!D0&1`9$!D0&0`$```%`0D!&`$``-at-``&`$``-at-``"0$``0``
M`"-at-``!-at-!&0&X;`#`HP`3`+Y```4``.-at-+ZA,2&W0BUBD``$D`2-at-!*`$D```!)
M`$D`2-at-!)````20!)`$H`20```$D`20!*`$D```!)`$D`2-at-!*`!(W#P``.`\`
M`$8/``!,#P``6`\``&0/``!E#P``9-at-\``'(/``"!#P``-at--at-\``(,/``"$#P``
MA0\``*`/``"A#P``H-at-\``*,/``"D#P``I0\``+-at-/``"Y#P``N-at-\``+L/``"\
M#P``O0\``,T/``#.#P``SP\``-`/``#E#P``Y-at-\```00```%$```!A````L0
M```,$```UP`!F`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8
M!_``UP`!F`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_``
MUP`!F`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_``UP`!
MF`?P`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_``UP`!F`?P
M`-0``;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#+``&8!_``UP`!F`?P`-0`
M`;8*\`#/``&D!_``SP`!,`;P`,\``2(%\`#/``&8!_```````````P``"0$8
M`0`$```%`-at-D!&`$``-at-``&`$`*```&`$9`;AL`,"C`!,`OD``!0``C-at-L*%!(;
M#"%\*0``20!*`$D`20```$D`20!)`$D```!)`$D`20!)````20!)`$D`20``
M`$D`20!)`$H`)`P0```-$```)!```"40```F$```-1```#80```W$```3!``
M`$T0``!.$```3Q```%`0``!1$```;1```&X0``!O$``` {!```'$0``!R$```
MC!```(T0``".$```CQ```)`0``"1$```JQ```*P0``"M$```KA```*\0``"P
M$```RA```,L0``#,$```S1```- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&
M\`#+``$B!?``SP`!F`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/
M``$B!?``SP`!F`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/``$B
M!?``SP`!F`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/``$B!?``
MSP`!F`?P`- {``9-at-'\`#4``&V"O``QP`!I`?P`,\``3`&\`#/``$B!?``SP`!
MF`?P`- {``9-at-'\`#4``&V"O``SP`!I`?P`,\``3`&\`#/``$B!?``````````
M`````P``!0(8`0`#```)`1-at-!``0```4""0$8`0`"```8`0`H```8`1D!N&P`
MP*,`$P"^0``%``"."PH4$AL,(7PI``!)`$H`20!)````20!)`$D`20```$D`
M20!)`$D```!)`$D`20!)````20!)`$D`2-at-`CS1```,X0``#/$```Z1```.H0
M``#K$```[!```.T0``#N$```"1$```H1```+$0``#!$```T1```.$0``*A$`
M`"L1```L$0``+1$``"X1```O$0``3!$``$T1``!.$0``3Q$``%`1``!1$0``
M9A$``&T1``!N$0``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&
M\`#[``$B!?``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\`#[
M``$B!?``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\`#[``$B
M!?``^P`!F`?P`-(``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\`#[``$B!?``
M^P`!F`?P`*8``9-at-'\`#/``&V"O``^P`!I`?P`/L``3`&\``````````````H
M```8`1D!N&P`P*,`$P"^0``%``"."PH4$AL,(7PI``!)`$H```!)````20!)
M````20```$D`20```$D```!)`$D```!)````20!)````2-at-```-at-``&`$`*```
M&`$9`;AL`,"C`!,`OD``!0``C-at-L*%!(;#"%\*0``20!*`$D`20```$D`20!)
M`$D```!)`$D`20!)````20!)`$D`20```$D`20!)`$H```0```4""0$8`1UN
M$0``;Q$``'`1``!Q$0`` {A$``(41``"&$0``XQ$``/L``2(%\`#[``&8!_``
MT-at-`!F`?P`-```3`J\`#0``$P*O``T``!,"KP`-```3`J\```````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M````````````````````````````````````````````````````````````
M``````````````$````H```8`1D!N&P`P*,`$P"^0``%``"."PH4$AL,(7PI
M``!)`$H`2-at-!)````20!)`$H`20```$D`20!*`$D```!)`$D`2-at-!)````20!)
M`$H`2-at-``!```!0()`1-at-!!PX`#P`(``$`2P`/`````P` {``!`\?\"`!P`!DYO
M {FUA;``"````!-at-!A"01C&``````````````````````````B`$%`\O^A`"(`
M%D1E9F%U;'0-at-4&%R86=R87!H($9O;G0```````````````````#C#-at-``!0#C
M$0``!0#K$0```P`$(?__`0``(?__`-at-`$(/__`P``````[P4``+T+``#C#-at-``
M```!`````0`!`````-at-````````,``& {2```*```#```%!```_`0```-at-&``"+
M!-at-``/P {````(``#!"```9PD``#`*```8"P``R`L``%L,```^#0``$`X``+L.
M```W#P``#!```,T0``!N$0``XQ$```L`#``-``X`#P`0`!$`$-at-`3`!0`%0`6
M`! {`&``9`!H`&P` {`!T`'-at-##`-at-Y! {G1H=7(-at-4BX-at-2&%N9"Y-86-I;G1O {V-at--at-
M2$0Z5&5M {&]R87)Y($ET96US.E=O {F0-at-5V]R:R!&:6QE($$-at-#D%R=&AU {B!2
M+B!(86YD%4%22"`Y-CI&964-at-0V]M {&%R:7-O;-at-Y! {G1H=7(-at-4BX-at-2&%N9$)-
M86-I;G1O {V-at--at-2$0Z36EC {F]S;V9T($]F9FEC93I-:6-R;W-O9G0-at-5V]R9"`V
M.D-%348Z1F5E($-O;7!A {FES;VX.07)T:'5R(%(N($AA;F0O36%C:6YT;W-H
M($A$.E1E;7!O {F%R} 2!)=&5M {SI7;W)D(%=O {FL-at-1FEL92!!(#,.07)T:'5R
M(%(N($AA;F0O36%C:6YT;W-H($A$.E1E;7!O {F%R} 2!)=&5M {SI7;W)D(%=O
M {FL-at-1FEL92!!(#,.07)T:'5R(%(N($AA;F0O36%C:6YT;W-H($A$.E1E;7!O
M {F%R} 2!)=&5M {SI7;W)D(%=O {FL-at-1FEL92!!(#,.07)T:'5R(%(N($AA;F1"
M36%C:6YT;W-H($A$.DUI8W)O {V]F="!/9F9I8V4Z36EC {F]S;V9T(%=O {F0-at-
M-CI#14U&.D9E92!#;VUP87)I {V]N#D%R=&AU {B!2+B!(86YD0DUA8VEN=&]S
M:"!(1#I-:6-R;W-O9G0-at-3V9F:6-E.DUI8W)O {V]F="!7;W)D(#8Z0T5-1CI&
M964-at-0V]M {&%R:7-O;-at-Y! {G1H=7(-at-4BX-at-2&%N9$9-86-I;G1O {V-at--at-2$0Z36EC
M {F]S;V9T($]F9FEC93I-:6-R;W-O9G0-at-5V]R9"`V.D-%348Z1F5E($-O;7!A
M {FES;VXM36%C#D%R=&AU {B!2+B!(86YD14UA8VEN=&]S:"!(1#I-:6-R;W-O
M9G0-at-3V9F:6-E.DUI8W)O {V]F="!7;W)D(#8Z0T5-1CI&964-at-0V]M {&%R:7-O
M;BU00_]``X`!````````````"``!``````````(``````````0`!`*\"U-at-$]
M`!4&``!-`%1I;65S($YE=R!2;VUA;-at-`,!-at-```-at-!3} 6UB;VP`"P8``$T`07)I
M86P`"P:0`4T`5&EM97,`(-at-`$`#$(C!-at-``-`"``!H`0````!OX-at-.&;^(#A-at-``
M```"``````````````````````````0`-at-Q``````````````````````````
M`````````"$#`````$H````:0T]-4$%225-/3B!/1B!%32!54T52($9%15,`
M```.07)T:'5R(%(N($AA;F0.07)T:'5R(%(N($AA;F0`````````````T,\1
MX*&Q&N$``````````````````````#L``__^``D`!-at-`````````````````!
M`````0```````!```````-at-````'____^``````````#_________________
M____________________________________________________________
M____________________________________________________________
M____________________________________________________________
6________________________________
`
end


--=====================_19263778==_
Content-Type: text/plain; charset="US-ASCII"


--=====================_19263778==_--




From: Bob Citron :      Bob_Citron-at-cc.chiron.com
Date: 3/28/96 8:37 AM
Subject: Elmers glue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello - is there anybody out there who can tell me where
to obtain 'Elmers glue'.

Bo


_____________________________________________________________________
Bo Johansen E-Mail: BoJ-at-bot.ku.dk
Botanical Institute Vioce: +45 3532 2157
Gothersgade 140 FAX: +45 3313 9104
DK-1123 Copenhagen K, Denmark http://www.bot.ku.dk/www/staff/boj.htm
---------------------------------------------------------------------







From: Cheri Owen :      cowen-at-cmb.biosci.wayne.edu
Date: Thu, 28 Mar 1996 10:32:05 -0500 (EST)
Subject: Re: freezing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In response to your question for freezing brain tissue. We freeze
brain for cryostat sectioning in our lab. After it is perfused,
(generally in 4% paraformaldehyde) we run the brain through a sucrose
series of 10%, 20% and 30% solution. This is for cryoprotection. We
then make a cup with aluminum foil, place a small amount of Tissue Tek
(freezing medium) in the bottom and place our 2mm slab of whole brain in
the cup with the face we want to section, down. We then submerge the
rest of the slab in Tissue tek and place the cup on dry ice with acetone
poured over the dry ice to make it colder. It freezes pretty rapidly,
and will last in the -20' freezer for a couple of weeks before sectioning.

Let me know if you want more details...

Cheri Owen
Detroit Neurotrauma Institute
Wayne State University
Detroit, MI
313-577-4648

On 27 Mar 1996, Igor Polyakov wrote:

} Subject: Time:9:42 AM
} OFFICE MEMO freezing Date:3/26/96
}
} I am looking for available information about freezing a tissue(brain) for LM
} histochemistry and immunocytochemistry. Also, I need information how to
} estimate the optical density of a histochemically stained sections? Are there
} books, those could give me such information?
} Any information would be very much appreciated.
} Igor_Polyakov-at-qmgate.arc.nasa.gov
}
}
}




From: ENERGY BEAM SCIENCES, INC :      75767.640-at-CompuServe.COM
Date: Thu, 28 Mar 1996 15:45:52 +0000
Subject: Polaron targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: lfox-at-wpo.it.luc.edu

I am pleased that your Polaron unit continues to give service. The
target life of a system is entirely dependant on the use and there is
not a standard figure anyone can give.

When deterioration of the target is significant as you describe, it
is time to replace it. No real damage will be done to the system,
but the quality will deteriorate and any contamination from behind the
target will be exposed to the plasma.

If you have a medium to high res. SEM or where grain islands from the
gold becomes a problem, then we recommend gold/palladium (SC502-314B)
or platinum (SC502-314C)

Your local agent EBS will be able to offer delivery and price.

If you experience any problems, please contact me.

Best regards
Tony King

Regards,

Tony King
Product specialist
VG Microtech/ Polaron range

Tel: +44 (0)1825 746251
Fax: +44 (0)1825 768343

Disclaimer:
The views and opinions expressed are not necessarily
those of Fisons plc or VG Microtech.







From: Joe D Geller :      geller-at-world.std.com
Date: Thu, 28 Mar 1996 11:38:41 -0500 (EST)
Subject: Calibration Standards

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



1) For typical calibration intervals for recertification you must check with
NIST and get their recommendation in writing for your auditors satisfaction. We specify one year intervals for our NIST traceable magnification
reference standards.

2) For NIST recertification services I remember one experience last year where NIST suggested that the purchase price of a new standard (SRM-484) would be comparable to recertifying an existing one. That, combined
with an out of stock situation left the customer with no solution.

3) For recertification of a magnification scale standard it is necessary to
have an "independent source" determine if there has been any physical or
dimensional degradation of the standard which would raise questions about its
usefulness. When we recertify standards they are cleaned (using several
different methods) to remove physical debris including deposited oils from
scanning electron microscopes and re-measure them.

4) EDX standards are used for both energy and intensity calibration. I would
suggest that comment #2 above again applies. Please note that NIST has very
few appropriate (homogeneous to the micrometer level) standards. They include
some glasses (SRM 1872-3) and the Cu-Au (SRM-482) and Au-Ag (SRM-481) series.
NIST also has a publication (NBS 26028) on the proper preparation techniques
for SRM 481-2). The general point about traceable standards is that the
composition be known to a certain accuracy. As stated in ISO guide 25 the
accuracy (and measurement precision) must be better than needed for your
analysis.


Joe Geller
Geller MicroAnalytical Laboratory
426e Boston St.
Topfield, MA 01983-1212
508 887-7000, fax: 887-6671
geller-at-tiac.com

-Offering analytical services (SEM, EPMA, Auger, metallography, profilometry),
standards (magnification, elemental, compound, alloy, glasses, and minerals)
for microanalysis, and EPMA computer control systems (WDS and EDS).




From: paulc-at-arms.gps.caltech.edu (Paul K. Carpenter)
Date: Thu, 28 Mar 1996 10:33:05 -0800
Subject: Re: EMPA: anybody ever done hair?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02110101ad808afb1a99-at-[131.215.67.110]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

John Fournelle wrote:

} Does anyone out there have any experience with EMPA of hair? Preparation,
} mounting, etc? Someone here is interested in looking at heavy metals that
} thru metabolism are concentrated in the hair.The levels will be low, so
} the MDL will be a question, but I don't know how the hair will stand up to
} an electron beam.

If you have enough material (hair, that is) it seems that XRF is a good
method. You could simply see what elements generate peaks, and then set up
for quantitative analysis. I suppose you could actually chop up the hair
samples to make a pressed XRF disk for quant analysis...

If you only have a few hairs and they must be analyzed as is, then EPMA is
going to be tough. You could try micro-XRF using a Mo thin foil mounted
above the sample; same comments as above for bulk XRF.

Paul


+------------------------------------------------------------+
| Paul K. Carpenter |
| Division Analytical Facility |
| Geological and Planetary Sciences MC 170-25 |
| California Institute of Technology |
| Pasadena, CA 91125 |
| 818-395-6126 (X-ray Lab) 818-568-0935 (FAX, Departmental) |
| paulc-at-arms.gps.caltech.edu |
+------------------------------------------------------------+






From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Thu, 28 Mar 1996 16:11:32 -0500 (EST)
Subject: Re: WORKING WITH PARAFFINIZED SECTIONS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Thu, 28 Mar 1996, Stanley Hayes wrote:

} I WOULD LIKE SOME INPUT FROM ANYONE WITH METHODS FOR DEALING WITH
} PARAFFINIZED SECTIONS NOW ON SLIDES AND TREATED FOR IN SITU
} HYBRIDIZATION AS WELL AS UNPARAFFINIZED SECTIONS. THE INVESTIGATOR
} WISHES TO IDENTIFY BY TEM THE TYPE OF CELLS LABELED. THANK YOU IN
} ADVANCE.
}
} S FRED. HAYES
}
Stan:

What I've done in the past for paraffinized sections is as follows:

Put slides in coplan jar and place in 60C oven............10 min.
Remove from oven and placr in xylene.......................3 min
xylene.......................3 min
100% ETOH....................3 min
95% ETOH....................3 min
80% ETOH....................3 min
70% ETOH....................5 min
50% ETOH....................5 min
DH2O...................2 X 10 min
2.5% Glut in PBS...........60 min
PBS....................2 X 15 min
1% OsO4 in PBS.............30 min
PBS....................3 X 15 min
50% ETOH...................10 min
70% ETOH...................10 min
95% ETOH...................10 min
100% ETOH...............3 X 10 min
Propylene Oxide..........2 X 10 min
1:1 PO:embedding resin..........60 min
1:2 PO:embedding resin..........60 min
1:3 PO:embedding resin..........60 min
100% embedding resin..........60 min

Before curing, drain off as much resin as possible, leaving a thin film.
Place a cured block on top of the area you want and cure for the
recommended time and temp for your resin.
Take off of slide by putting the slide on dry ice for a few minutes then
snap the block off. Section should come off of slide.

Hope this helps.

Peace,

Phil Rutledge
8-{)





From: Charles A. Garber :      103532.3325-at-CompuServe.COM
Date: 28 Mar 96 13:17:27 EST
Subject: Gold sputter targets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On March 27, the following was posted:


Seeing Linda Fox's post just reminded me of something I found out the other
day here. Apparently the cheapest source of gold (99.9% pure) is new 'limited
edition' gold coins. Don't ask me why! One guy here had a crown about an inch
across which he then rolled to produce a 2 inch target for an industrial
coater. Saved the company a packet.

Richard Beanland
GMMTL Caswell,
Towcester,
Northants NN12 8EQ.

Tel. +44 1327 356363
Fax. +44 1327 356775
Email richard.beanland.gecm.com
==================================

For the almost twenty years that our firm has been supplying replacement sputter
coater gold targets, we have used the very highest gold starting purities
available, namely "four nines" or 99.99% pure gold.

Now I must start to wonder again, as I did twenty years ago, is such purity
really "necessary"? Clearly a lower level purity would be able to be offered at
a cheaper price.

The consensus that I seem to detect is that purity is very much on the minds of
the typical purchaser of a sputter coater cathode. The user wants reproducible
results and does not want to have stray elements present in the event EDS work
should be done on the now gold coated samples.

Also, as the impurity levels increase, assuming they are not precious metals but
are base metals and other elements, since such elements don't really sputter at
the typical voltages used in the sputter coaters typically found in an EM
laboratory, it is presumed that there is a "build up" of such impurity elements
on the target surface, reducing the sputtering rate, thereby increasing the
coating time, and exposing
the sample to far more electron and radiant heating than would otherwise be the
case.

Now I find it hard to believe that the typical user, with extremely costly
instrumentation, would be willing to take such risks to "save" such a small
monetary amount.

One reason why at least some purchased sputter coater cathodes appear expensive
is that the manufacturers supply them only as 3 mil or 5 mil thick targets. The
"fabrication" charge is the same whether for a thicker or thinner target. That
is why the SPI cathodes are all 10 mils thick, they last a lot longer and the
"fabrication" cost is a smaller percentage of the total cost or making the
target.

Disclosure: SPI Supplies manufactures gold and other replacement
targets for most manufacturers of sputter coaters and we would have an
obvious vested interest is not having SEM users rolling limited edition gold
coins into sputter coater targets!

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service: SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: em-at-mediacity.com (Ed Monberg)
Date: Thu, 28 Mar 1996 23:02:35 -0700
Subject: Re: Calibration Standards

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01540b02ad7fb91dc884-at-[205.216.172.10]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Yes, Good Morning !




An ISO 9000 slogan (to put on Watson's Forehaead):



Please rest assured that "ISO9000"

Implementation

will ensure that if you happen to make a poor product,

your product will be poor consistently.






Regards,



(signed) Ed Monberg {em-at-mediacity.com}

--------------------------------------------------

510-429-1060 Fax 429-1065
LMDC, (Laser Motion Development Co.)
3101 Whipple Road
Union City, CA 94587-1216


For our Most recent Catalogue of "On Hand" EQUIPMENT:
Send empty mail to: {Cat-at-lasermotion.com}


Our web page: http://www.lasermotion.com (Is beginning to take shape!)
Our e-mail: office-at-lasermotion.com

{-------------------------------- Our page width
-----------------------------}






From: nano-at-ns1.LIHTI.ORG (Nanoprobes Inc.)
Date: Fri, 29 Mar 1996 07:11:10 -0500
Subject: Osmium tetroxide / IGSS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Everybody:

I've received several comments recently about the fate of silver-enhanced
gold particles (in EM) when specimens are treated with osmium tetroxide.
The literature indicates that this process can cause a slight reduction in
the size of the particles, probably due to re-oxidation of the deposited
silver. However, the effect seems to be very variable: in some cases many
of the particles disappear, or the color fades from the sections. Has
anyone had problems with this, and has anyone found a good way to prevent
it?

Thanks,

Rick Powell

*****************************************************************
* NANOPROBES, Incorporated *
* 25 East Loop Road, Suite 124, Stony Brook, NY 11790-3350, USA *
* Tel: (516) 444-8815 Fax: (516) 444-8816 *
* *
* http://www.tiac.net/users/everlast/nano *
*****************************************************************







From: Bo Johansen :      BOJ-at-bot.ku.dk
Date: Fri, 29 Mar 1996 15:17:42 GMT+0100
Subject: Thanks!!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This list i amazing - I've got 32 answers to my question about Elmers glue.
Thanks to all of you.

Bo

_____________________________________________________________________
Bo Johansen E-Mail: BoJ-at-bot.ku.dk
Botanical Institute Vioce: +45 3532 2157
Gothersgade 140 FAX: +45 3313 9104
DK-1123 Copenhagen K, Denmark http://www.bot.ku.dk/www/staff/boj.htm
---------------------------------------------------------------------






From: chuck (c.p.t.) o'dale :      odale-at-bnr.ca
Date: Fri, 29 Mar 1996 09:18:00 -0500
Subject: SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X400-Received:
by mta bnr.ca in /PRMD=BNR/ADMD=TELECOM.CANADA/C=CA/; Relayed; Fri, 29 Mar 1996 09:43:12 -0500
X400-Received:
by /PRMD=BNR/ADMD=TELECOM.CANADA/C=CA/; Relayed; Fri, 29 Mar 1996 09:18:38 -0500
X400-Received:
by /PRMD=BNR/ADMD=TELECOM.CANADA/C=CA/; Relayed; Fri, 29 Mar 1996 09:18:00 -0500
[/PRMD=BNR/ADMD=TELECOM.CANADA/C=CA/;bcars520.b.416:29.02.96.14.18.38]
X400-Content-Type: P2-1984 (2)
Content-Identifier: SEM

I received this address off the net RE: SEM.

Is this a news/exchange group?

My interests are in E-Beam testing, using SEM to probe functioning
integrated circuits to extract temporal and voltage information.

Chuck




From: Scott Holt :      102467.2752-at-CompuServe.COM
Date: 29 Mar 96 10:05:43 EST
Subject: TEM Epoxy: EPOTECH 353ND

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

After the recent discussion of TEM epoxies, John Mardinly wrote to me requesting
any information on the toxic properties of Gatan's G1 Epoxy, as Gatan's MSDS
sheet did not mention any carcinogens. Since this is really Epoxy Technology's
EPOTECH 353ND, I will give the information I have regarding all-in-one.

Part B (hardener) of the 353ND product contains 0.05wt% of Acrylonitrile {CAS#
107-13-1, Toxicity Data: ACGIH(TLV-SKIN): TWA=2ppm}. Acrylonitrile appears on
the Mass Substance List. According to OSHA, acrylonitrile is a suspected cancer
causing agent.

The reason there is no indication of acrylonitrile on most MSDS sheets regarding
this product is that the quantities are below required reporting levels.
Imidazole is the only other hazardous ingredient in Part B of the epoxy, and
this should be reported on the MSDS due to significant levels.

For a copy of Epoxy Technology's MSDS sheet, please call them at (508)663-9782.

Hope this helps.

Best regards,
BUEHLER
41 Waukegan Rd.
Lake Bluff, IL 60044
(847)295-4546

BUEHLER WWW Page: http://www.buehlerltd.com






From: Alan Brooker :      alanb-at-jeolsys.demon.co.uk
Date: Fri, 29 Mar 1996 16:19:06 0000
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please subscribe me!

Alan
*---------------------------------------*----------------------------------*
| Dr.Alan Brooker | Email: alanb-at-jeolsys.demon.co.uk |
| Product Support Manager | |
| JEOL (UK) Ltd., JEOL House, | Tel: +44 (0) 707 377117 |
| Watchmead, Welwyn Garden City, | Fax: +44 (0) 707 373254 |
| Herts, AL7 1LT, United Kingdom | |
*---------------------------------------*----------------------------------*





From: ayache-at-csnsm.in2p3.fr (Ayache Jeanne)
Date: Fri, 29 Mar 1996 18:49:21 +0000
Subject: inscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603291832.SAA129053-at-smtp-gw01.ny.us.ibm.net}

Dear colleague

I would like to belong to the users of the microscopy server.and receive
informations from microscopists.
I have a recent paper on preparation sample for TEM without ion milling :

Application of the ionless tripod polisher to the preparation of YBCO
superconducting multilayers and bulk ceramics thin films.
J.ayache and P.H Albar=E8de, Ultramicroscopy 60 (1995) 195-206

Best regards

Dr Jeanne Ayache

Centre de Spectroscopie Nucleaire et Spectrometrie de Masse
CSNSM -CNRS
Batiment 108
91405 Orsay Campus
=46rance
Tel :33 1 69 41 52 19
fax :33 1 69 41 52 68
Email : ayache -at-csn-hp.in2p3.fr






From: kna101-at-utdallas.edu
Date: Fri, 29 Mar 1996 08:41:37 -0600 (CST)
Subject: Re: WORKING WITH PARAFFINIZED SECTIONS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: allowed
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Company-Confidential

In my experience, the fixation quality of sections prepared for paraffin,
not to mention in situ hybrid., is too pour to get much informtion of
type of cell at EM level. But, I have reembedded tissue after removing
the media. The slides the tissue were on had been coated with gelatin
before the tissue was placed on them. I filled half a gelatin capsule with
new media and turned it on end, on the slide. Once the media had set up,
I pried the capsule off the slide. I could section the specimen this
way, but as I said the quality of the tissue was poor for EM. Good luck.

Karen Pawlowski

On Thu, 28 Mar 1996, Stanley Hayes wrote:

} I WOULD LIKE SOME INPUT FROM ANYONE WITH METHODS FOR DEALING WITH
} PARAFFINIZED SECTIONS NOW ON SLIDES AND TREATED FOR IN SITU
} HYBRIDIZATION AS WELL AS UNPARAFFINIZED SECTIONS. THE INVESTIGATOR
} WISHES TO IDENTIFY BY TEM THE TYPE OF CELLS LABELED. THANK YOU IN
} ADVANCE.
}
} S FRED. HAYES
}




From: Bruce Cutler :      BCutler-at-eureka.chem.ukans.edu
Date: Fri, 29 Mar 1996 15:47:26 -0500 (CDT)
Subject: Re: WORKING WITH PARAFFINIZED SECTIONS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: Raymond F Egerton :      egerton-at-phys.ualberta.ca
Date: Fri, 29 Mar 1996 14:44:39 -0700 (MST)
Subject: Mike Whelan festschrift

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



There will be a special issue of MICRON to honour Mike Whelan, one of the
pioneers of transmission electron microscopy in the physical sciences.

It will contain review and original-research papers written by scientists
who have worked in TEM imaging, defect analysis, electron diffraction and
electron energy-loss spectroscopy.

Manuscripts are due in September of this year. For further information,
contact Ray Egerton (egerton-at-phys.ualberta.ca).





From: Corvos-at-aol.com
Date: Fri, 29 Mar 1996 18:05:01 -0500
Subject: Re: Material Safety Data Sheets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Scott Walck,

Every supplier must send a MSDS with all shippments under the DOT
regulations. And they must notify the shipping agent of the contents of every
box including the US postal service.

As an example: Even the smallest amount of oil, even in a pump it has to have
a MSDS when shipping. If the pump leaks and their is not a reference on the
out side of the package, in some states you must call in a HAZ-MAT team for
clean-up. It can be a very expencive package.

Walter Protheroe







From: Edmund Glaser :      eglaser-at-umabnet.ab.umd.edu
Date: Fri, 29 Mar 1996 10:40:47 -0500 (EST)
Subject: Re: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Tue, 26 Mar 1996 scott_l-at-unin1.unorth.ac.za wrote:

} Are there IBM compatible PC based programmes able to assist with doing
} stereology from TEM and LM micrographs. What are the hardware require-
} ments?
}
} Thank you
} Leon Scott
} SA
}

You should look into the recently released design based
stereology software, called Stereo Investigator. It's from
MicroBrightField in Colchester,VT, USA. Stereo Investigator
employs optical or physical disectors in conjunction with the
optical fractionator. This permits analysis of serial sections or
micrographs or live microscope images. The software includes 2D and 3D
mapping capabilities. It also lets you perform programmed scans of video
or optical images. The hardware requirements are basically those of a PC
with a video camera. MBF will give you specifics.

You can reach MicroBrightField at 802-655-9360 or by email:
info-at-microbrightfield.com.

They also have a web site: http://www.microbrightfield.com/microb

Edmund Glaser, D. Eng.
Dept. Physiol.
Univ. Md. School. Med.
Baltimore, MD 21201 USA
Ph: (410) 706-5041
Fax: (410) 706-8341





From: klivi-at-jhu.edu (Kenneth JT Livi)
Date: Fri, 29 Mar 1996 10:39:50 -0400 (EDT)
Subject: Re: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,

Wil's recent comment on the safety hazards of distilled water brought to
mind some peculiar safety regulations here in MD. In reference to liquid N
(which can be dangerous stuff), we were first required to wear gloves while
handling the stuff. Then came goggles and soon after that, full face
shields. The funny thing is, the most dangerous aspect of our handling of
LN2 is that students often wear sandals in the summer and are very likely
to get stung by droplets. There are no safety measures for feet
protection! But now that I have mentioned it, some occupational safety
officer listening in will recommend new safety procedures requiring
protective booties!
In the end, we can't legislate common sense, nor can we abdicate
responsibility to those above.

(The opinions above are mine and of anyone who agrees with them.)

Kenneth JT Livi
Department of Earth and Planetary Sciences
34th and Charles Streets
The Johns Hopkins University
Baltimore, Maryland 21218
(410) 516-8342 (voice)
(410) 516-7933 (fax)
klivi-at-jhu.edu (e-mail)






From: M.V. Parthasarathy :      mvp2-at-cornell.edu
Date: Fri, 29 Mar 1996 18:16:26 -0400 (EDT)
Subject: RE: Looking for the reference in full by Luft J.H., 1966

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

In message Fri, 29 Mar 96 19:31:57 -0800,
"Alberto J. Villena" {villena-at-ibm.net} writes:

} Dear colleagues,
}
} I'm lookink for the following reference in full:
}
} } Luft J.H., 1966: Fed. Proc, 25 1773
----------------

Here is the full reference:

Luft, J.H. 1966. Fine structure of capillary and endocapillary layer as
revealed by ruthenium red. Fed. Proc., 25:1773.

*************************************************************************
M.V. Parthasarathy
Professor of Plant Biology &
Director, Cornell Integrated Microscopy Center
Section of Plant Biology
228 Plant Science Building
Cornell University, Ithaca, NY 14850
Telephone: (607) 255-1734
Fax: (607) 255-5407
E-mail: mvp2-at-cornell.edu
***********************************************************************




From: Scott Holt :      102467.2752-at-CompuServe.COM
Date: 29 Mar 96 13:54:03 EST
Subject: TEM Epoxy: EPOTECH 353ND

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

After the recent discussion of TEM epoxies, John Mardinly wrote to me requesting
any information on the toxic properties of Gatan's G1 Epoxy, as Gatan's MSDS
sheet did not mention any carcinogens. Since this is really Epoxy Technology's
EPOTECH 353ND, I will give the information I have regarding all-in-one.

Part B (hardener) of the 353ND product contains 0.05wt% of Acrylonitrile {CAS#
107-13-1, Toxicity Data: ACGIH(TLV-SKIN): TWA=2ppm}. Acrylonitrile appears on
the Mass Substance List. According to OSHA, acrylonitrile is a suspected cancer
causing agent.

The reason there is no indication of acrylonitrile on most MSDS sheets regarding
this product is that the quantities are below required reporting levels.
Imidazole is the only other hazardous ingredient in Part B of the epoxy, and
this should be reported on the MSDS due to significant levels.

For a copy of Epoxy Technology's MSDS sheet, please call them at (508)663-9782.

Hope this helps.

Best regards,
BUEHLER
41 Waukegan Rd.
Lake Bluff, IL 60044
(847)295-4546





From: Daniel L. Callahan :      dlc-at-owlnet.rice.edu
Date: Fri, 29 Mar 1996 15:37:15 -0600 (CST)
Subject: epoxies and confidentiality

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Good Afternoon:

A large number of microscopists in the hard sciences and electronics
industry use epoxies like Epoxy Technology 353ND or Gatan G-1 (which may
or may not be the same thing). Since epoxies are basically commodities,
these companies have a vested interest in not revealing their sources and
compositions. However, as a physical scientist, sometimes it behooves me
to know the chemicals I am using in the experiment. Basically, I would
like to know the composition of Epo Tech 353ND sufficiently to be able to
determine potential sample interactions and health risks without
revealing sufficient information to readily reproduce the product. I
don't know if this is possible: it may be that anyone knowing the secret
ingredients can easily determine their proportion and processing. On the
other hand, it may be that any competent polymer chemist could guess the
composition reasonably accurately.

I would particularly appreciate commentary from the business providers
who watch the list. Incidentally, if you do know the composition please
do NOT post it directly to the list unless you know that it has been
published or patented (which is also public documentation). This is
someone's business interest.


Daniel L. Callahan
Department of Mechanical Engg. and Materials Science
Rice University
dlc-at-owlnet.rice.edu




From: Dr. L. P. Stoter :      LPS-at-teknesis.demon.co.uk
Date: Sat, 30 Mar 1996 07:57:02 +0000
Subject: Re: Safety & Liquid N2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: (Unverified)
Message-Id: {v01510104ad82961fa812-at-[158.152.199.245]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Wil's recent comment on the safety hazards of distilled water brought to
} mind some peculiar safety regulations here in MD. In reference to liquid N
....
} officer listening in will recommend new safety procedures requiring
} protective booties!
} In the end, we can't legislate common sense, nor can we abdicate
} responsibility to those above.

Try getting your safety officer to conduct an experiment:

1. Hold out hand,
2. Pour a small volume of liquid N2 over hand
3. Now the interesting bit - put on a glove, and pour the same quantity of
liquid N2 into glove.
4. Phone for ambulance.

The point is that a brief contact causes no problems, but if the contact is
continued you get a nasty burn.

Gloves, goggles, masks (and shoes) are actualy more dangerous when handling
liquid N2 than sandles and no protection. And clothes are actually more
dangerous than being naked. Get the safety officer to experiment. With a
little persuasion you can probably convince the safety officer that when
handling liquid N2, everbody should be naked.

More seriously, bureaucrats, administrators and the inexperienced should
talk to somebody who has real knowledge.

--------------------------------------------------------------
Dr. Larry Stoter
Technesis
17, Rocks Park Road, Uckfield, E. Sussex, TN22 2AT, United Kingdom
Larry-at-teknesis.demon.co.uk
--------------------------------------------------------------






From: Finn-Mogens Haug :      f.m.s.haug-at-basalmed.uio.no
Date: Sat, 30 Mar 1996 12:02:32 +0100
Subject: Re: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603301051.LAA15533-at-pons.uio.no}
X-Sender: finnmog-at-pons.uio.no
X-Mailer: Windows Eudora Pro Version 2.1.2
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

On Tue, 26 Mar 1996 scott_l-at-unin1.unorth.ac.za wrote:
}
} Are there IBM compatible PC based programmes able to assist with doing
} stereology from TEM and LM micrographs. What are the hardware require-
} ments?
}
} Thank you
} Leon Scott
} SA
}

I did not see anyone mention:

Kinetic Imaging Ltd.,
South Harrington Bldg
Sefton Street
Liverpool L3 4BQ
UK
Tel +44(0) 151 709 8661
Fax +44(0) 151 709 8633
E-mail: kineticimage-at-aol.com

They offer "Digital Stereology", running under DOS/Windows-3, HW requirements
486/33 MHz, 8MB RAM, 200MB Hard disk, High resolution monitor.
Imaging board by Matrox or Univision. CCD-camera. It is not quite clear
whether you
also need to purchase their "Fenestra" image processing package.

I suppose many would wish a stereology subsystem to behave as an add-on to
whater
image processing system (IPS) he or she is already used to. This might take
the form of
an "autonomous" system running on a PC with only a high resolution graphics
display board
and sufficiently processing power and RAM (and capable of dealing with
TIFF-images).
Or it might take the form of a function library which could integrated into
any IPS capable
of linking dll libraries (such as Image Pro Plus, Optimas, analySIS,
Visilog, to mention a
few IPS's which I believe have this capability).

So far the range of stereology software seems to be much smaller than that
of general
image analysis systems. However, with the increasing use of digital images,
there
should be room for many such products. Any more feedback to the original
query????

Best regards
*****************************************************************
Finn-Mogens Haug
University of Oslo, Institute of basic medical sciences,
Department of anatomy E-mail: f.m.haug-at-basalmed.uio.no
Box 1105 Blindern Phone : +47 22 85 12 67
N-0317 Oslo, NORWAY Fax : +47 22 85 12 78
*****************************************************************






From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Sat, 30 Mar 1996 19:00:21 -0600
Subject: Re: LN2 safety (was no subject)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Dear Microscopists,
}
} Wil's recent comment on the safety hazards of distilled water brought to
} mind some peculiar safety regulations here in MD. In reference to liquid N
} (which can be dangerous stuff), we were first required to wear gloves while
} handling the stuff. Then came goggles and soon after that, full face
} shields. The funny thing is, the most dangerous aspect of our handling of
} LN2 is that students often wear sandals in the summer and are very likely
} to get stung by droplets. There are no safety measures for feet
} protection! But now that I have mentioned it, some occupational safety
} officer listening in will recommend new safety procedures requiring
} protective booties!
} In the end, we can't legislate common sense, nor can we abdicate
} responsibility to those above.
}
} (The opinions above are mine and of anyone who agrees with them.)
}
} Kenneth JT Livi

Kenneth:
Your students are correct, it's the safety officers who are in
error. Any good safety manual for cryogenic gases specifically recommends
sandals, and *against* shoes, gloves, and the like.
The reason is the Leidenfrost effect. The LN2 droplets
flash-evaporate when they hit bare skin, producing an insulating layer of
gas, and prevents freezing damage. Assuming it's a quick spill, and not
continous contact. Shoes, glove (including asbestos gloves), and any other
clothing that confines the LN2, and cold gas *will* cause freezing.
A face shield is a good measure, byt the gloves and any other
similar regs are dangerous.
Phil

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: em-at-mediacity.com (Ed Monberg)
Date: Sat, 30 Mar 1996 22:38:16 -0700
Subject: Re: LN2 safety (was no subject)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01540b14ad83c6ff379c-at-[205.216.172.10]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

As a wild and fun loving farm boy who went into science, I found LN2 to be
a wonderful toy. The closest I ever came to danger was when an associate
put some in a nalgene bottle and closed the top, resulting in a frozen
plastic flask going "POP".

We did stunts along the lines of freezing the contents of our associates
lunch pails, shattering a fish (boy was that an ultimately smelly mistake
!), and once when we froze a rubber tube in the neck of a dewar by mistake,
we had to empty the entire 50 liters into a trash container.

Consider the heat of evaporation, compare it to the heat of condensation of
steam, and think how many burns YOU personally know of from that source vs.
from LN2.

I am repelled at the lack of savy and the excess of fear instilled by the
law ("profession") regarding such pseudo threats as LN2.

The presumptive (and ignorant) reaction is irrational, but accfording to
the same principles, TWO safety officers should come in the box containing
each microwave oven, a gun should bear a judge attached, and an auto, a
member of the supreme court, or a receipt for $1000 from your favorite
lobbyist. Remember, not only can the law be bought, It's really CHEAP !


!!!!!!


Good luck. Just act normal.





Regards,



(signed) Ed Monberg {em-at-mediacity.com}

--------------------------------------------------

510-429-1060 Fax 429-1065
LMDC, (Laser Motion Development Co.)
3101 Whipple Road
Union City, CA 94587-1216


For our Most recent Catalogue of "On Hand" EQUIPMENT:
Send empty mail to: {Cat-at-lasermotion.com}


Our web page: http://www.lasermotion.com (Is beginning to take shape!)
Our e-mail: office-at-lasermotion.com

{-------------------------------- Our page width
-----------------------------}






From: Robert.R.Wise-at-Sparc5.Microscopy.Com
Date: Sun, 31 Mar 1996 11:02:41 +0000
Subject: modems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Although not strictly a microscopy question, the members of this
list seem to know a lot about computers as well. So here goes...

Does anyone have any recommendations on the proper modem for
connecting a Macintosh (Centris 610 with EtherNet card) at home to a local
phone line? US Robotics has a 28.8 kbps model for about $200. What does
the 28.8 stand for? Is this just a plug in or do I have to get something
like an adapter (such as an Asante' FriendlyNet Media Adapter)? Please
help the computer illiterate.

Bob Wise






From: zaluzec-at-Microscopy.Com (Nestor J. Zaluzec)
Date: Sun, 31 Mar 1996 12:41:20 -0600
Subject: Submission of Papers for the 96 MSA/MAS/MSC-SMC Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



SUBJECT: Submission of Papers for the 96 MSA/MAS/MSC-SMC Meeting

********************************
* *
* MICROSCOPY & MICROANALYSIS -96 *
* Joint Annual Meeting of *
* MSA/MAS/MSC-SMC *
* *
********************************


SPECIAL EXTENSION FOR RECEIPT OF ABSTRACTS


Some microscopists and microanalyists who wished to present papers at
Microscopy & Microanalysis '96 did not receive their Registration Bulletin /
Call for Papers package in time to submit by the published deadline (March
15). Because of this, a SPECIAL ARRANGEMENT has been made to allow papers
to still be accepted until April 10th 1996 (receipt of manuscript at the
address listed below). If you wish to submit but have not, please contact
the Business Office (BusinessOffice-at-MSA.Microscopy.Com 1-800-538-3672)
as soon as possible, for any additional details or questions on when and
where to send
your paper.

You may download all of the submission information INCLUDING forms
and instructions from the Web at:

http://www.msa.microscopy.com


At this late date, please mail copies directly as follows:


Original & 1 Copy:

Microscopy & Microanalysis-96
4 Barlows Landing Rd.
Suite 8
Pocasset, Ma. 02559 USA.



2 Copies:

Dr. Nestor J. Zaluzec
Microscopy & Microanalysis-96
797 Bonnie Brae Ct.
Bolingbrook, Illinois 60440 USA



-----------------------------------------------------------------------







From: rw9-at-psu.edu (Rosemary A. Walsh)
Date: Sun, 31 Mar 1996 21:24:23 GMT
Subject: Re: EM User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy-at-Sparc5.Microscopy.Com (MSA Listserver)

At 9:52 AM 3/28/96 -0500, Hand,Arthur wrote:
} Since there have been a large number of requests for the data on user fees, I
} have decided to post the information on the listserver as attachments to
} this message. They were created in MS Word 6; one attachment is in Mac
} format; the other is in PC format. If you have problems opening them, let me
} know and I will fax the data to you as soon as possible.
}
} Arthur R. Hand
} Central EM Facility
} UConn Health
} Ctr
}
} Content-Type: application/octet-stream; name="Fee Comparison-Mac"
} Content-Transfer-Encoding: x-uuencode
}
} Attachment converted: MacintoshHD:Fee Comparison-Mac (????/----) (0000215E)
} Content-Type: application/octet-stream; name="Fee Comparison-PC"
} Content-Transfer-Encoding: x-uuencode
}
} Attachment converted: MacintoshHD:Fee Comparison-PC (????/----) (0000215F)


Arthur,
Thanks for your time and effort--I had a problem opening the Mac file.
Would you please fax me the file? My FAX number is 814-863-1357.

Rosemary Walsh
EM Facility
The Biotechnology Institute
519 Wartik Lab
The Pennsylvania State University
University Park, PA 16802






From: Robert.R.Wise-at-Sparc5.Microscopy.Com
Date: Sun, 31 Mar 1996 15:56:57 +0000
Subject: LN Dewar

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I need to purchase a 10 liter Dewar for LN transfer and storage. Does
anyone have any recommendations for an appropriate product? I would like
to find one with rate of boil-off of less than one liter per day (such
specs are rarely given in catalogs).

Thanks in advance,

Bob


Robert R. Wise, PhD
Director, UWO Electron Microscope Facility
Department of Biology
University of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404 tel
(414) 424-1101 fax
wise-at-vaxa.cis.uwosh.edu






From: Finn-Mogens Haug :      f.m.s.haug-at-basalmed.uio.no
Date: Sun, 31 Mar 1996 22:28:10 +0100
Subject: Re Stereology (re-posting)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199603312117.XAA22651-at-pons.uio.no}
X-Sender: finnmog-at-pons.uio.no
X-Mailer: Windows Eudora Pro Version 2.1.2
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I take the liberty of re-posting this message, which had
a terrible layout. Also, the e-mail address has been verified
and seems to work.

On Tue, 26 Mar 1996 scott_l-at-unin1.unorth.ac.za wrote:
}
} Are there IBM compatible PC based programmes able to assist with doing
} stereology from TEM and LM micrographs. What are the hardware require-
} ments?
}
} Thank you
} Leon Scott
} SA
}

I did not see anyone mention:

Kinetic Imaging Ltd.,
South Harrington Bldg
Sefton Street
Liverpool L3 4BQ
UK
Tel +44(0) 151 709 8661
Fax +44(0) 151 709 8633
E-mail: kinetimage-at-aol.com (or Kinetimage-at-aol.com)

They offer "Digital Stereology", running under DOS/
Windows-3, HW requirements 486/33 MHz, 8MB RAM, 200MB
Hard disk, High resolution monitor. Imaging board by
Matrox or Univision. CCD-camera. It is not quite clear
whether you also need to purchase their "Fenestra" image
processing package.

I suppose many would wish a stereology subsystem to behave
as an add-on to whatever image processing system (IPS) he
or she is already used to.

It might take the form of an "autonomous" system running
on a PC with only a high resolution graphics display board
and sufficient processing power and RAM (capable of dealing
with TIFF-images). Or a function library which could
be integrated into any IPS capable of linking dll libraries
(such as Image Pro Plus, Optimas, analySIS, Visilog, to
mention some which I believe have this capability).

With the increasing use of digital images, there should be
room for many such products. Any more feedback to the
original query?

Best regards






******************************************************************
Finn-Mogens Haug
University of Oslo, Institute of basic medical sciences,
Department of anatomy E-mail: f.m.haug-at-basalmed.uio.no
Box 1105 Blindern Phone : +47 22 85 12 67
N-0317 Oslo, NORWAY Fax : +47 22 85 12 78







From: Hand-at-nso1.uchc.edu (Hand,Arthur)
Date: Thu, 28 Mar 1996 09:52:35 -0500
Subject: EM User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

by msgate.uchc.edu (PostalUnion/SMTP(tm) v2.1.8d for Windows NT(tm))
id AA-1996Mar28.095100.1274.161755; Thu, 28 Mar 1996 09:52:35 -0500
Received: from msgate.uchc.edu (msgate.uchc.edu [155.37.9.14]) by Sparc5.Microsc
opy.Com (8.6.11/8.6.11) with ESMTP id IAA07569 for {Microscopy-at-MSA.Microscopy.Co
m} ; Thu, 28 Mar 1996 08:52:59 -0600
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11) id IA
A07572 for dist-Microscopy; Thu, 28 Mar 1996 08:53:02 -0600
Received: from Sparc5.Microscopy.Com (sparc5.microscopy.com [206.69.208.10]) by
ormail.intel.com (8.7.4/8.7.3) with SMTP id AAA28130; Sun, 31 Mar 1996 00:48:57
-0800 (PST)
Received: from ormail.intel.com by relay.hf.intel.com with smtp
(Smail3.1.28.1 #2) id m0u3Ion-000qDLC; Sun, 31 Mar 96 00:49 PST




From: Hand-at-nso1.uchc.edu (Hand,Arthur)
Date: Thu, 28 Mar 1996 09:52:35 -0500
Subject: EM User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

by msgate.uchc.edu (PostalUnion/SMTP(tm) v2.1.8d for Windows NT(tm))
id AA-1996Mar28.095100.1274.161755; Thu, 28 Mar 1996 09:52:35 -0500
Received: from msgate.uchc.edu (msgate.uchc.edu [155.37.9.14]) by Sparc5.Microsc
opy.Com (8.6.11/8.6.11) with ESMTP id IAA07569 for {Microscopy-at-MSA.Microscopy.Co
m} ; Thu, 28 Mar 1996 08:52:59 -0600
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11) id IA
A07572 for dist-Microscopy; Thu, 28 Mar 1996 08:53:02 -0600
Received: from Sparc5.Microscopy.Com (sparc5.microscopy.com [206.69.208.10]) by
ormail.intel.com (8.7.4/8.7.3) with SMTP id AAA28130; Sun, 31 Mar 1996 00:48:57
-0800 (PST)
Received: from ormail.intel.com by relay.hf.intel.com with smtp
(Smail3.1.28.1 #2) id m0u3Ion-000qDLC; Sun, 31 Mar 96 00:49 PST




MicroscopyListserver Archive Email Extraction Software Version NJZ07060908