From: ENERGY BEAM SCIENCES, INC :      75767.640-at-CompuServe.COM
Date: 01 Apr 96 09:16:15 EST
Subject: Re: re: safety in EM labs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Ken,
You're right about not being able to legislate common sense. In the case of
liquid nitrogen, I have seen LN2 slopping out of styrofoam cups onto peoples'
clothes, into instruments, all over papers, etc.
For all of the examples everyone can cite of *unnecessary* regulations, I am
still amazed to see kitchen microwave ovens in labs being used to heat toxic
chemicals, including fixatives and heavy metals, the fumes from which then
"ventilate" into the open lab when the door is opened. In some cases, the oven
is then used to heat coffee and/or lunch.
Steven Slap
75767,640-at-compuserve.com





From: William Tivol :      tivol-at-wadsworth.org
Date: Mon, 1 Apr 1996 11:02:49 -0500 (EST)
Subject: Re: Safety & Liquid N2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Gloves, goggles, masks (and shoes) are actualy more dangerous when handling
} liquid N2 than sandles and no protection.

True with the exception that the shutoff valve handle on a big
LN2 tank will get cold enough to be dangerous, and does not exhibit the
leidenfrost effect. Using a glove or other insulation when turning off
the LN2 after filling the dewar makes good sense.
}
} More seriously, bureaucrats, administrators and the inexperienced should
} talk to somebody who has real knowledge.
}
A great general rule. Reading the manual also helps, and every
lab should have a safety manual on hand.
Yours,
Bill Tivol




From: Kovacs Arpad :      femkov-at-gold.uni-miskolc.hu
Date: Mon, 1 Apr 1996 17:39:29 GMT
Subject: Re: Safety & Liquid N2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html







From: Weiming Yu :      weiming-at-LFD.physics.uiuc.edu
Date: Mon, 1 Apr 1996 12:15:03 -0600 (CST)
Subject: Temperature controller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friends:

Does anyone know which company sells temperature controller for optical
microscope (specially for old immersion objective) ?
Thank you.

Weiming Yu, Ph.D.
Dept. of Physics, UIUC
1110 W. Green
Urbana, IL 61801
weiming-at-lfd.physics.uiuc.edu





From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Mon, 1 Apr 1996 13:16:16 -0500
Subject: HRP substrates

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d04ad85ccdb2569-at-[128.206.15.189]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I plan to return to some HRP-based histochemistry and thought I
would stir up some controversy by asking what you all feel the best
precipitating substrate for immunocytochemistry with peroxidase labels. I
have always used DAB without metals. What is the disadvantage of using
metal intensification? Does anybody have experience with the "stable"
solution of DAB (e.g., Pierce Chemicals Metal Enhanced DAB Substrate Kit)?
Are they worth the convenience? TIA.


Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: Dave King (607)757-1248 T37/257-3 Ext DEKING-at-VNET.IBM.COM
Date: 1 Apr 1996 10:08:33 EST
Subject: Re: EM User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604011506.JAA12875-at-Sparc5.Microscopy.Com}

To: Microscopy ListSer {Microscopy-at-MSA.Microscopy.Com}
*** Reply to note of 04/01/96 07:19

We do outside analytical work over the complete spectrum of
materials, chem lab and product assurance areas. We charge
different rates for each instrument and basically different
operations. Sample prep, data analysis, and report writing get
charged, as well as actual SEM time, for example.

We figure it as "beginning to end." Think about how much time
you'd have saved if you didn't have that job at all. Everything
must be covered, or someone else is subsidizing it.

. {
} { {
} } ===========} Dave King { { {
} } } ------------------------------------------------------ { { { {




From: mlamvik-at-mcnc.org
Date: Mon, 1 Apr 1996 16:28:54 -0500
Subject: Getting data from Kevex 8000?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know how to get spectrum data out of a Kevex 8000 analyzer into
a desktop computer for further analysis? Advice will be appreciated.

Thanks,
Michael Lamvik
MCNC, Research Triangle Park, North Carolina






From: Richard Thrift :      Richard_Thrift-at-depotech.com
Date: Mon, 01 Apr 1996 11:44:05 -0800
Subject: Re: microscope safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s15fc16b.056-at-depotech.com}
X-Mailer: Novell GroupWise 4.1

I have to tell you about this: after reading the mail on this topic I used the
far door to enter our microscope room. I rarely use this door. It has a
sign (which I've never noticed before; it must have been there for the last
half-year) saying "NOTICE Safety glasses required in this area"

Richard

} } } Kenneth JT Livi {klivi-at-jhu.edu} 03/29/96 06:39am } } }
Dear Microscopists,

Wil's recent comment on the safety hazards of distilled water brought to
mind some peculiar safety regulations here in MD. In reference to liquid N
(which can be dangerous stuff), we were first required to wear gloves
while handling the stuff. Then came goggles and soon after that, full face
shields. The funny thing is, the most dangerous aspect of our handling of
LN2 is that students often wear sandals in the summer and are very likely
to get stung by droplets. There are no safety measures for feet
protection! But now that I have mentioned it, some occupational safety
officer listening in will recommend new safety procedures requiring
protective booties!
In the end, we can't legislate common sense, nor can we abdicate
responsibility to those above.

(The opinions above are mine and of anyone who agrees with them.)

Kenneth JT Livi
Department of Earth and Planetary Sciences
34th and Charles Streets
The Johns Hopkins University
Baltimore, Maryland 21218
(410) 516-8342 (voice)
(410) 516-7933 (fax) klivi-at-jhu.edu (e-mail)








From: Warren Straszheim :      wesaia-at-iastate.edu
Date: Mon, 01 Apr 1996 09:12:22 -0600
Subject: Re: EM User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We had momentary problems with the files on a Mac. We first had to locate
them, our Eudora didn't tell us which folder it dropped it in. It ended up
in SYSTEM\EUDORA FOLDER\ATTACHMENTS, which we should have guessed.

We could not simply double-click on it to open it. Apparently it did not
have the extra information to tell the desktop what kind of file it was.
Therefore, we had to first open up Word, then open the document the
old-fashioned way through the FILE/OPEN menu. I have found that this is
often required and usually works for documents that come through different
channels.

Being a Windows man myself, I am used to opening up Word and leaving it
running throughout the day so it is a simple matter to switch to Word and
open a new document through the pull-down menus. Sometimes the normal
associations get lost or naming conventions are not followed. But the
pull-dopwns almost always work.

At 09:24 PM 3/31/96 +0000, you wrote:
} Arthur,
} Thanks for your time and effort--I had a problem opening the Mac file.
} Would you please fax me the file? My FAX number is 814-863-1357.
}
} Rosemary Walsh
} EM Facility
} The Biotechnology Institute
} 519 Wartik Lab
} The Pennsylvania State University
} University Park, PA 16802
----------------------------------------------------
Warren E. Straszheim
270 Metals Development, Ames Lab/ISU, Ames IA, 50011
Phone: 515-294-8187 FAX: 515-294-3091

E-Mail: wes-at-ameslab.gov (or: wesaia-at-iastate.edu)
http://www.public.iastate.edu/~iprt_info/cfce/ (re: coal)
http://www.public.iastate.edu/~wesaia/marl/ (re: SEM)

coal characterization and processing
electron microscopy, x-ray analysis, image analysis
computer applications





From: Yves Maniette :      yves-at-giga.sct.ub.es
Date: Mon, 1 Apr 1996 16:29:16 +0000 (GMT)
Subject: Re: C2 movement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy List {Microscopy-at-Sparc5.Microscopy.Com}


To Jan van der Meijden

thank you for your mail,

I just checked what you recommend. The displacement is about 20 mm, so
within your specifications. Despite it is within Philips specification, this
does not allow to work easily from bright field-dark field and
diffraction modes, because when working at about 100 000 times (usual
magnification rate in order to make BF DF diff images), the movement
comes to be 20*20= 400 mm, which is over my own specifications (i.e. the
light goes out of the screen therefore I have to get back to 20 000
magnification in order to see where it is...).

The only reasonable way I have found is to press the intensity limit
softkey when being near the cross over in order to prevent from going too
low in intensity of the C2 lens. This works, but I will welcome with
great pleasure another method to get rid of the problem.

On the other hand, during the alignment procedure, due to this shift
problem, we do not know exactly how to center correcty the light. I am
used to press at each step the normalization button (i.e. photo button),
but am amazed because after the whole process (canon alignment) usually
the light does NOT get back in the center of the screen. Well maybe it is
within specifications, I haven't checked this out.

Thank you for every new input, I am sure that many Philips microscope users
are eager to know more about the subject. For this reason I an sending a
copy to the list as well.

Sincerely,


Yves MANIETTE
Universitat de Barcelona
Serveis Cientifico Tecnics
Unitat ESCA TEM
Carrer Luis Sole i Sabaris
E-08028 BARCELONA

Tel +34 3 402 16 95
Fax +34 3 402 13 98

On Thu, 28 Mar 1996, Jan van der Meijden wrote:

} Dear Sir
}
} Via a detour, your request to know the C2 movement when changing C2
} has arrived on my desk.
} The procedure we use in the factory is as follows:
} Magnification 5800x
} Focus C2 and centre the beam.
} Fully over focus the beam and focus again. Centre the beam again
} Fully underfocus the beam and focus again. Measure the distance from
} the centre. This distance should be less than 5cm.
} According to your mail this distance on your scope is about 7cm at
} 20.000x and thus far within specification.
} The problem is caused by the high remanence in the C2 lens and the C2
} lens not perfectly mechanically aligned in the optical axis.
} To overcome the problem use one spotsize higher or lower in order to
} do a light normalisation.
}
} Regards Jan van der Meijden
}
}
}
}
} Yours sincerely
}
}
}
} Jan van der Meijden
}




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Tue, 2 Apr 1996 13:20:10 +1100
Subject: EM: Marine invertebrate texts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: st004718-at-brandywine.otago.ac.nz
Message-Id: {v01540b00ad863cf89037-at-[139.80.120.140]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear Microscopists,
Does anyone out there know of a textbook or any other reference material
which covers integumental ultrastructure - cuticle, epithelium and
associated structures and also oogenesis - in marine invertebrates. We have
a marine science student who is studying subclass Ascothoracida who is
having problems getting the ultrastructure info she needs.

Thanks in advance.

Richard


Richard Easingwood
South Campus Electron Microscope Unit
School of Medical Sciences
University of Otago
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

SOUTHERNMOST E.M UNIT IN THE WORLD















From: rh208-at-cus.cam.ac.uk (Ray Hicks)
Date: Mon, 1 Apr 1996 15:47:11 +0000
Subject: Re: Microscope Incubators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: rh208-at-pop.cus.cam.ac.uk
Message-Id: {v01530501ad8587d712d3-at-[131.111.80.78]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi Brian,
I think that most CO2 sensors work by measuring the thermal conductivity or
indirectly the heat capacity of the gas mix, this makes them awkward to use
in a situation where relative humidity and temperature may fluctuate (since
both of these will affect the temperature of the sensor). They also might
(I don't know) need a large volume of gas to pass over them to register a
change.
I would suggest that you might look into three alternative strategies

i) the simplest is to use a cannister of premixed gas, presuming that you
want to have a constant pCO2.

ii)Monitor the pH of the medium colorimetrically, assuming the colour
change of the medium is mainly due to dissolved CO2, which may not be the
case.

iii) Measure the infra-red absorbance of the gas mix, I believe that this
method is used in some incubators, but I'm not sure what wavelength to use.


(i) is the easiest and cheapest assuming you don't want to vary the pCO2,
if you did you could adapt it by using two bottles, one of CO2 and one of
air, these methods might cause humidity regulation problems. (ii) would
have to be verified, and if it worked you'd have to build your own probe
(I'd have a go if you like), but you would have the benefit of regulating
dissolved CO2 not just gaseous. Depending on the absorption spectrum of
CO2, it might be hard to find suitable detectors and emitters for (iii).

I hope this helps

Ray

At 2:13 pm 29/3/96, Brian Burgess wrote:
} Hi Everyone
} We are designing an incubator to fit on a microscope to set the
} temperature at 37C and control humidity and maintain a level of
} CO2 at 5%. I have had a real hard time locating a CO2
} control/sensing unit. I did find one marketed by Forma
} Scientific but it is lunky and quite large. I rather need a
} smaller unit. As mentioned before I need to control temperature
} and humidity levels. Does anyone know of a good system or a
} manufacturer of such a system?

} ----------------------------------------------
} Brian Burgess
} Chemical Engineering
} University of Florida
} e-mail: burgbr-at-che.ufl.edu

Ray Hicks
________________________________________________________________________
|University of Cambridge |Tel 01223 330149 |
|Department of Medicine |Fax 01223 336846 |
|Level 5, Addenbrookes Hospital |e-mail rh208-at-cus.cam.ac.uk |
|Hills Road Cambridge |Web Page/ facsmac.med.cam.ac.uk |
|CB2 |ftp server 131.111.80.78 |
|UK | |
|_________________________________|_____________________________________|






From: !Microscopy-request-at-Sparc5.Microscopy.Com (Robert Kayton,MAC,CROET)
Date: Mon Mar 18 12:38:43 -0500 1996
Subject: Re: Microscope Incubators

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: kayton-at-ohsu.edu (Robert Kayton)
Message-Version: 2
} To: microscopy-at-Sparc5.Microscopy.Com (!Microscopy)

Title: Biological Science Technician (Plants)
Lab: USDA-ARS, Salinas, CA

The USDA-ARS is seeking a biological science technician (plants)
(GS-404-7, 8, or 9) for the Crop Improvement and Protection Research Unit
in Salinas, CA. The incumbent will share responsibilities in electron
microscopy of plant virus infections for ultrastructural characteristics.
The incumbent will also assist in research involving molecular, serological,
and biological studies of several plant viruses infecting sugarbeet and
vegetables. Candidate must have a knowledge of electron microscopy,
plant virology, and knowledge of microbiological techniques. Must be a U.S.
citizen. Bachelors degree is desirable. Salary is commensurate with
experience ($24,610-39,140 per annum).
For information regarding research program contact Gail C. Wisler or
James E. Duffus (408)755-2835. For information regarding application
procedures/forms contact Tom Nelson (408)755-2810. Applications must be
postmarked by May 6, 1996. The USDA is an equal opportunity employer.





From: ENERGY BEAM SCIENCES, INC :      75767.640-at-CompuServe.COM
Date: 01 Apr 96 09:16:50 EST
Subject: Re: LN2 Dewar

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bob-
Energy Beam Sciences offers a range of stainless steel LN2 lab containers, from
5L up to 220L. The 10L container has a loss rate of .8 liters per day. Please
contact me directly by e-mail or telephone (800-992-9037) for part numbers,
prices, etc.
Steven E. Slap, Vice-President
75767,640-at-compuserve.com





From: kna101-at-utdallas.edu
Date: Mon, 1 Apr 1996 07:55:16 -0600 (CST)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Kenneth:

I understand your frustration with all the regulations that just keep
coming for hazardous material in the work place, but I have to point out
the flip side of this issue. I know of a lab that recently got into
trouble because for years the regulators were leaving it up to the head
of the lab to make sure precautions were taken. This lab was solely made
up of students and the professor/lab head. The professor was leaving
everything up to the students to do and learn, giving little if any
dirrection to them. These students came from accounting and psycology
and english-major backgrounds. They had no experience working in an
environment full of hazardous substances, such as a histology lab. They
were working with flourescent microscopy, without using the proper
filters and hazardous substances without the proper protection, ie
gloves, fume hoods. All waste was going down the drain! This is an
EXTREME example, I know. I just wanted to show you how bad it can get if
some sort of overseeing isn't in place.

I've worked in histology labs
for 16 years and I do get tired of having the rules changing about once a
year, ie- we have a built-in book self near a fire door. It's been there
3-4 years. Last month the inspectors came through and told us to take it
down because it's too close to the door. They've inspected that bookcase
every year and not until now was it a problem! So I do know how you feel.

Karen Pawlowski

On Fri, 29 Mar 1996, Kenneth JT Livi wrote:

} Dear Microscopists,
}
} Wil's recent comment on the safety hazards of distilled water brought to
} mind some peculiar safety regulations here in MD. In reference to liquid N
} (which can be dangerous stuff), we were first required to wear gloves while
} handling the stuff. Then came goggles and soon after that, full face
} shields. The funny thing is, the most dangerous aspect of our handling of
} LN2 is that students often wear sandals in the summer and are very likely
} to get stung by droplets. There are no safety measures for feet
} protection! But now that I have mentioned it, some occupational safety
} officer listening in will recommend new safety procedures requiring
} protective booties!
} In the end, we can't legislate common sense, nor can we abdicate
} responsibility to those above.
}
} (The opinions above are mine and of anyone who agrees with them.)
}
} Kenneth JT Livi
} Department of Earth and Planetary Sciences
} 34th and Charles Streets
} The Johns Hopkins University
} Baltimore, Maryland 21218
} (410) 516-8342 (voice)
} (410) 516-7933 (fax)
} klivi-at-jhu.edu (e-mail)
}
}
}




From: ldm2-at-apollo.numis.nwu.edu (L.D.Marks)
Date: Mon, 1 Apr 1996 05:57:14 -0600
Subject: Re: EM User Fees

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy-at-Sparc5.Microscopy.Com

I have a general question about user fees, in particular how various people
bill for hours. We use a "gentle" system with a computer interfaced to the
microscope to track real hours, rather than a harsh system where one bills
from the minute that the user walks in the door. As a consequence, we have
comparatively low hours/week, and I have recently been criticized for this.
I would be very interested in comments both from users and administrators
about this - should TEM facilities bill like lawyers or not?




From: ychen-at-MACC.WISC.EDU
Date: Sun, 31 Mar 1996 23:33:15 -0600
Subject: Re: modems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Date: Sun, 31 Mar 1996 11:02:41 +0000
} From: Robert.R.Wise-at-Sparc5.Microscopy.Com
} Subject: modems
} To: Microscopy-at-Sparc5.Microscopy.Com
} X-Sender: wise-at-vaxa.cis.uwosh.edu
} MIME-version: 1.0
} Content-type: text/plain; charset="us-ascii"
} Content-transfer-encoding: 7BIT
}
} Although not strictly a microscopy question, the members of this
} list seem to know a lot about computers as well. So here goes...
}
} Does anyone have any recommendations on the proper modem for
} connecting a Macintosh (Centris 610 with EtherNet card) at home to a local
} phone line? US Robotics has a 28.8 kbps model for about $200. What does
} the 28.8 stand for? Is this just a plug in or do I have to get something
} like an adapter (such as an Asante' FriendlyNet Media Adapter)? Please
} help the computer illiterate.
}
} Bob Wise


Bob,
The speed of a modem is measured by Bit per Second (bps). The 28.8kbps
modem will double the transmission speed than the 14.4 kbps one. The 28.8k
modem uses error correction and data compression protocols called v.42bis,
MNP2-5 for high speed data transmission. The installation is simple: just
plug the modem cable (suppose this is a Mac modem) in the serial port
marked as MODEM on your Mac, and connect phone line. You need to set up or
configurate your communication software before start. EtherNet card is
used for network connection, not for modem. If you have any question,
please contact me. Good luck.
Ya Chen


Ya Chen

Integrated Microscopy Resource (IMR)-- III M M RRRRRR

an NIH Biomedical Research Resource I M M M M R R
University of Wisconsin, Madison, WI I M M M RRRRRR
1675 Observatory Drive #167 I M M R R
Madison, WI 53706 I M M R R
TEL : 608-263-8481 I M M R R

FAX : 608-265-4076 III M M R R

Email:YChen-at-macc.wisc.edu
IMR WWW Home Page: http://www.bocklabs.wisc.edu/imr.html
2nd Symposium on Integrated Microscopy: Sept. 20-22, 1996






From: !Microscopy-request-at-Sparc5.Microscopy.Com (Sverker Enestrom)
Date: Tue Mar 19 09:13:29 +0100 1996
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: kayton-at-ohsu.edu (Robert Kayton)
Message-Version: 2
} To: Microscopy-at-Sparc5.Microscopy.Com


} This information is at least 4 years old and may not be correct.
} About 4 years ago (maybe a little longer) there was a lengthy review
} article in either New England Journal of Medicine or JAMA regarding
} the slow human neurological disorders. At that time, there were at
} least 5 recognized diseases that were classified as "Alzheimer" or
} "Alzheimer-like". One of these was suspected to be of prion origin,
} although by now it may be classified as something else.
}
} One thing to consider if you are involved in studies involving human
} (or any mammalian) brain tissue is that in years past, prion diseases
} were commonly misdiagnosed as Alzheimers, and as humans (even
} pathologists) are not perfect, this still may be the case.

AD and prion diseases share the presence of amyloid precursor proteins
(beta-PP, APrP) but they are distinct diseases, both starting at the synapse
(beta-PP and PrP are proteins of the neuromuscular junction and CNS synapse).
In some of the prion diseases prion amyloid plaques are seen together with
paired helical filaments (NFT) making them "Alzheimer-like" as Budy writes.


*********************************************************
Sverker Enestrom M.D., Ph.D.
Department of Pathology
University of Linkoping, Sweden
Phone: +46 13 22 15 20
Fax: +46 13 13 22 57
*********************************************************






From: !Microscopy-request-at-Sparc5.Microscopy.Com (Robert Kayton,MAC,CROET)
Date: Mon Mar 18 16:40:07 EST 1996
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: kayton-at-ohsu.edu (Robert Kayton)
Message-Version: 2
} To: Microscopy-at-Sparc5.Microscopy.Com

} As far as I know Alzheimers disease is not infectous---- Where did you get your
} information about Alzheimers disease?

This information is at least 4 years old and may not be correct.
About 4 years ago (maybe a little longer) there was a lengthy review
article in either New England Journal of Medicine or JAMA regarding
the slow human neurological disorders. At that time, there were at
least 5 recognized diseases that were classified as "Alzheimer" or
"Alzheimer-like". One of these was suspected to be of prion origin,
although by now it may be classified as something else.

One thing to consider if you are involved in studies involving human
(or any mammalian) brain tissue is that in years past, prion diseases
were commonly misdiagnosed as Alzheimers, and as humans (even
pathologists) are not perfect, this still may be the case.



W. L. Steffens, Ph.D
Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: !Microscopy-request-at-Sparc5.Microscopy.Com (Robert Kayton,MAC,CROET)
Date: Mon Mar 18 10:04:59 EST 1996
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: kayton-at-ohsu.edu (Robert Kayton)
Message-Version: 2
} To: Microscopy-at-Sparc5.Microscopy.Com

} Has anyone heard of the slow viruses that can survive gluteraldehyde fixation?

The infectious agents that are able to survive fixation are the
prions. These agents contain protein only...no nucleic acid. Among
these are Creutzfeld-Jakob, scrapie, kuru,
Gersten-Straussler-Schenker, BSE (mad cow disease) and apparently
some forms of Alzheimers.

It is known that formaldehyde does not inactivate them, not surprising
since formaldehyde is one of the least cross-linking of all
fixatives, so far as proteins are concerned. My understanding is
that glut. does inactivate them, but don't take my word for it.

} On a similar note, there was an MSA-sponsored speaker a couple of years ago who
} was encouraging EM labs to make money by offering virus identification sevices
} to medical centers. The preparation protocol included advice to accept unfixed
} (ie. infectious), unidentified material and prepare negative stained samples on
} the bench in the laboratory.

We are a CLIA and State certified human clinical laboratory, licensed for
negative-stain virus identification. This has always been part of
our function as a veterinary EM lab, and for about 5 years we have
offered it as a service to the community. We do make a considerable
amount of money from it. We accept samples only from certain
reference laboratories and practitioners and these are only stool
samples, mostly (} 95%) from infantile diarrheas. Our submission form has an
entry for any unusual precautions that must be taken (ie HIV,
hepatitis, etc). Our understanding with our clients is that we don't
process such samples.
W. L. Steffens, Ph.D



Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: Joiner Cartwright, Jr., Ph.D. :      joiner-at-bcm.tmc.edu
Date: Mon, 01 Apr 1996 17:09:46 -0600
Subject: Printers for digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fellow microscopists -

A few months ago there was a thread on the server dealing primarily with
flatbed scanners and the best technology for acquiring digital images. I may
have missed out on the part of the conversation dealing with printing
digital images, but perhaps one or some of you might summarize any thoughts
on what technologies were considered useful.

I run an EM lab in a Pathology department of a medical school. Most of my EM
is diagnostic work on kidney specimens and tumors. We printed out 4,500
images for each of the two past years. I would like to go digital, but
finding a printer that will handle this load cheaply and quickly, yet with
the required high quality is difficult. We plan to produce the usual 3 1/4"
X 4" negatives in our scope as we always have. Then we would scan the
negatives into a computer (IBM clone), storing them temporarily on a large
hard disk, and ultimately archiving them on a writable CD system. The prints
need not be durable as once they are studied and the case closed, they would
be trashed. The negatives would be archived as would be the image files on
the CD's.

The problem is that people are suggesting dye sub for quality and laser for
quick & dirty. I need QUICK & QUALITY......and cheap. Printing on a dye
sublimation printer would be too costly. There are some enhancement boards
that can be added to laser printers, and I have a friend who has printed
some of my scanned negatives on his Photoscan system, Philips's image
recording system. They come close to answering the problem, but I am
wondering what else there is out there. I believe that Philips's system may
be LaserPix or PhotoJet Plus from XLI Corp.

I realize that I may be asking you to repeat a conversation already carried
out, but if you could pass on the consensus opinions, I would appreciate it.


Joiner Cartwright, Jr., Ph.D.

Director, Electron Microscopy tel.: (713)798-4658
Department of Pathology, Rm.286-A FAX: (713)798-3945
Baylor College of Medicine Internet: joiner-at-bcm.tmc.edu
One Baylor Plaza Compuserve: 71555,1206
Houston, Texas 77030 U.S.A.






From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Mon, 01 Apr 1996 20:23:36 EST
Subject: Safety issues in EM lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

I know that my own safety committee gets under my own skin once in a
while.

But in over twenty six years of business, let me relate by far the most
serious injury that occurred in our firm. Considering that we work
around high voltage, use hazardous chemicals including osmium tetroxide,
deal with sometimes hazardous samples, what was our most serious
injury? Would you believe it involved Polaroid film?

One of the microscopists (who always claimed eye protection was not
needed in an SEM lab) was opening up a case of Type 52 film. The flaps
on the case, as it turned out were not creased, as they should have
been, therefore the tension present in the fold, when the sealing tape
on the top was slit with a razor, literally flew open, cutting the
corneal surface of one of the startled technician's eyes. Fortunately
sight was not lost in the eye, but there was a scar formed, and there
will be the need for lifetime care under a good ophthalmologist and who
knows what other impairments might show up with age.

Now in retrospect, nearly twenty years later, I still feel guilty about
not enforcing to an even greater degree, perhaps with the threat of
being dismissed, our own "rule" that safety glasses should be worn in
the laboratory. If an accident is going to happen, generally speaking
it will happen when it is least expected.

Now my point only is that some of these safety regulations, superfluous
as some of them might sound, often times do have some basis of logic
and rationale, my Polaroid film case being, I think, a good example.

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: !Microscopy-request-at-Sparc5.Microscopy.Com :      ron-anderson-at-vnet.ibm.com (1-914-892-2225)
Date: Thu Mar 28 10:09:47 EST 96
Subject: A little humor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: kayton-at-ohsu.edu (Robert Kayton)
Message-Version: 2
} To: MICROSCOPY-at-Sparc5.Microscopy.Com

Sorry, I couldn't resist sharing this... It was sent to me 'as is'
with no author. I'm certainly not clever enough to write this.

Ron


I am the Very Model of a Modern Teenage Cyberpunk*


I am the very model of a modern teenage Cyberpunk
I rent my own apartment and it's full of electronic junk
I own a VAX, a 486, I've even got a PDP
I've finished Myst and Doom but I am stumped by Wing Commander III

I'm very well aquainted too with matters pornographical
I have a list of image sites, both overseas and national
So if you want to see a picture of that Anna Nichole Smith
I'll fire up my terminal and fetch for you a naughty GIF

I'm totally an anarchist, the government I'd like to wreck,
Though if they were to get blown up, who'd give to me my welfare cheque?
In short if you need answers that concern your electronic junk,
I am the very model of a modern teenage Cyberpunk

I know the ancient myths about RTM, Pengo and Mitnick
I 'hack' into computers and I then perform a credit check
I scare all my non-hacker friends with tales of cracker theivery
and even though I'm spouting crap they'll listen and believe in me

I've learned to spot a troll and I've seen flames about the way I spell,
I've traced badly forged cancels and seen napalm poured on AOL
I've laughed at all the newbies and their flailing cries of "You all Suck!"
I've been flamed by Carasso, with an anvil I have then been struck

I've hung around in alt.tasteless and seen war waged on rec.pets.cats
I've spent my time in talk.bizarre and used those stupid Relay Chats
In short, if you need answers that concern your electronic junk,
I am the very model of a modern teenage Cyberpunk

Well postings like "MAKE.MONEY.FAST", I am now somewhat wary at,
I have been "Global Killfiled" by the Joel Furr Commissariat,
When rosebud posts a lengthy rant 'bout Microsoft she swears is true,
I know that she is just another short lived kook without a clue

When I have learnt what progress has been made upon the Internet,
When I know something more than just a smattering of netiquette,
In short when I can have a world-wide soapbox on which I can stand
I've got no time for other things, like beer and trips to Disneyland

My life outside the Internet is very very sad you see
I cannot get my spots to fade, my social life's a tragedy,
But still if you need answers that concern your electronic junk,
I am the very model of a modern teenage Cyberpunk.

(With apologies to G&S)





From: dietmar.reiter-at-uibk.ac.at (Dietmar Reiter)
Date: Tue, 2 Apr 1996 09:24:20 +0100
Subject: Re: EM: Marine invertebrate texts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: c71956-at-dm.uibk.ac.at
Message-Id: {v01530505ad8691f3ce30-at-[138.232.71.39]}
Mime-Version: 1.0
Content-Type: text/plain; charset=iso-8859-1
Content-Transfer-Encoding: quoted-printable

} Dear Microscopists,
} Does anyone out there know of a textbook or any other reference material
} which covers integumental ultrastructure - cuticle, epithelium and
} associated structures and also oogenesis - in marine invertebrates. We have
} a marine science student who is studying subclass Ascothoracida who is
} having problems getting the ultrastructure info she needs.

You might browse the "Microscopic Anatomy of Invertebrates" series
(Frederick W. Harrison, Burton J. Bogitsh, Eds.), Wiley-Liss, New York,
1991. It=B4s a highly comprehensive series of 15 volumes, covering in-depth
ultrastructure from Protozoa up to Hemichordata, Chaetognatha, and else.

-Dietmar-


+++ Dietmar Reiter {dietmar.reiter-at-uibk.ac.at} ++++++++++++++++
+++ Dept. of Zoology and Limnology, University of Innsbruck ++++
+++ Technikerstrasse 25, A - 6020 Innsbruck, Austria +++++++++
+++ ph: (+43)-512-507-6170 (-6161), fax: ..-507-2930 (-2957) +++






From: M.F. Butler :      mfb12-at-cus.cam.ac.uk
Date: Tue, 2 Apr 1996 13:24:48 +0100 (BST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



unsubscribe




From: David.Rothbard-at-ipst.edu (David Rothbard)
Date: Tue, 2 Apr 1996 07:57:15 -0500
Subject: RE: MSDS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {199604021254.HAA27062-at-IndyNet.indy.net}
To: "Joiner Cartwright, Jr., Ph.D." {joiner-at-bcm.tmc.edu} ,
Microscopy list {microscopy-at-Sparc5.Microscopy.Com}

DOT regulations DO NOT require that an MSDS accompany every shipment. DOT
does have strict requirements for packaging and labeling (exterior) that
are specific to the contents. DOT also requires that shippers of hazardous
material have a 24 hour telephone number (their own, or a service) that can
answer questions about the material. Both DOT and OSHA require training of
personel that handle hazardous materials.

David Rothbard

--
Institute of Paper Science and Technology






From: James R. Stets :      stetsjr-at-ttown.apci.com
Date: Tue, 2 Apr 1996 07:52:17 -0500 (EST)
Subject: Re: Kevex 8000 Data to PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Michael Lamvik asked about moving data from a Kevex 8000 to a PC. Kevex
used to sell a package to do this that cost about $1000. This bought you
a cable that connected to one of your printer ports on the 8000 and your
PC, a diskette with Kermit, software for the 8000, and a very thick manual.

I used it many times to move images from our 8000 to a Macintosh, and it
worked fine although it's not fast.

I don't know if Kevex sells this anymore since the 8000 is pretty old. A
gentleman named Robert Schaller was very helpful in getting us set up
with it, but this was at least 5 years ago and he may or may not still be
with Kevex.

Jim Stets
Air Products and Chemicals, Inc.
Allentown, PA
stetsjr-at-ttown.apci.com
My opinions are my own, not my employer's.




From: Schoonhoven, Robert :      rschoonh.sph-at-mhs.unc.edu
Date: Tue, 2 Apr 1996 11:23:58 -0500
Subject: Re: HRP substrates

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {F54A613101F70300-at-mhs.unc.edu}
In-Reply-To: {F422613101F70300}

We have been using the EnVision kit sold by DAKO with great success.
This kit does come with a "stable" DAB solution which works very well.
Under some conditions when we need to increase the intensity we use the
"Quick DAB Enhancer Solution" sold by Innovex Biosciences which works
very well and will not change the color of the DAB like nickle chloride.

The usual disclaimer applies to both of the companies mentioned above ie:
I'm simply a satisfied customer and recieve no renueration for naming
their products

Robert Schoonhoven
Laboratory of Molecular Carcinogenesis and Mutagenesis
University of North Carolina
CB#7400, Rosenau Hall
Chapel Hill, NC 27599-7400
Ph. 919-966-6343
Fax 919-966-6123





From: Joe D Geller :      geller-at-world.std.com
Date: Tue, 2 Apr 1996 08:52:31 -0500 (EST)
Subject: Re: SEM Image distortion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


If you have a digital imaging system it is possible to reduce the number
of sources of trouble. Does the distortion show up on the digitally
recorded image?

Joe Geller
Geller Microanalytical Lab
426e Boston St.
Topsfield, MA 01983-1212
508 887-7000

On Mon, 1 Apr 1996, Neuberger, Damian wrote:

}
} Hello All:
}
} We have a JEOL JSM-6300F microscope that exhibits a not-so-slight Y-image
} distortion. I hope that I can explain it clearly and that someone out there
} may have suggestion(s) as to the source of the problem that we have not been
} able to get resolved by field service personnel.
}
} Materials: (1) SEM; 2) NIST SRM 484f SEM magnification standard; 3) 2000
} copper mesh grid (Pella Cat #631C) that has been calibrated as a secondary
} standard with the NIST standard; 4) Polaroid Type 53 film, 4x5".
}
} Conditions: Image of the mesh grid recorded at X2000, 5 keV, 25 mm WD,
} aperture #4 (30 micrometers), probe current setting #8 (1x10-11), camera
} aperture setting 5.6, scan rate 80 (sec/scan?). Micrograph is oriented so
} that the 5" dimension is along the X-axis. Scan direction top to bottom
} (Y-axis). Scan rotation is OFF, tilt correction is OFF, specimen tilt is 0,
} SEI mode.
}
} Problem: The micrograph as well as the viewing monitors exhibit a
} distortion in the Y?-axis. That is, the vertical spacing between grid bars
} on one end of the X-axis of the photo is about 1.5 to 2 mm greater (out of
} 74mm) at the right end than on the left end. This distortion is easily seen
} in the image of the grid. (Wouldn't a photo in this email make things a
} whole lot easier!)
}
} Solutions: That's where you knowledgeable microscopists come in. Any ideas
} will be gratefully accepted and tested!
}
} Thanks so much, everyone.
}
} Damian Neuberger
} neuberd-at-baxter.com
}
}
}




From: Walt Bobrowski (313) 996-7814 :      bobroww-at-aa.wl.com
Date: Tue, 02 Apr 1996 08:11:53 -0400 (EDT)
Subject: Re: Printers for digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta SRVR05.MUAS; Relayed; Tue, 02 Apr 1996 08:11:53 -0400
Mr-Received: by mta SRVR05; Relayed; Tue, 02 Apr 1996 08:11:53 -0400
Mr-Received: by mta SRVR01; Relayed; Tue, 02 Apr 1996 08:13:54 -0400
Disclose-Recipients: prohibited
MSA MICROSCOPY MAILING LIST {MICROSCOPY-at-Sparc5.Microscopy.Com}
Message-Id: {3353110802041996/A12861/SRVR05/11A4120B3400*-at-MHS}
X-Envelope-To: joiner-at-bcm.tmc.edu, MICROSCOPY-at-MSA.MICROSCOPY.COM
Autoforwarded: false
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; CHARSET=US-ASCII
Content-Transfer-Encoding: 7BIT
Importance: normal
Priority: normal
Sensitivity: Company-Confidential
Ua-Content-Id: 11A4120B3400
X400-Mts-Identifier: [;3353110802041996/A12861/SRVR05]
Hop-Count: 2

Check out DOUBLE RES IV from Laser Printer Accessories Corp., a board you
install into your current laserjet II or III to get 600 dpi quality. We have
one and it really does make a BIG difference. 1-800-225-4098.

Walter F. Bobrowski
Subcellular Pathology
Parke-Davis Pharmaceutical Research
Ann Arbor, MI 48105

TEL: 313-996-7814
FAX: 313-996-5001
E-Mail: BOBROWW-at-AA.WL.COM


} Fellow microscopists -
}
} A few months ago there was a thread on the server dealing primarily with
} flatbed scanners and the best technology for acquiring digital images. I may
} have missed out on the part of the conversation dealing with printing
} digital images, but perhaps one or some of you might summarize any thoughts
} on what technologies were considered useful.
}
} I run an EM lab in a Pathology department of a medical school. Most of my EM
} is diagnostic work on kidney specimens and tumors. We printed out 4,500
} images for each of the two past years. I would like to go digital, but
} finding a printer that will handle this load cheaply and quickly, yet with
} the required high quality is difficult. We plan to produce the usual 3 1/4"
} X 4" negatives in our scope as we always have. Then we would scan the
} negatives into a computer (IBM clone), storing them temporarily on a large
} hard disk, and ultimately archiving them on a writable CD system. The prints
} need not be durable as once they are studied and the case closed, they would
} be trashed. The negatives would be archived as would be the image files on
} the CD's.
}
} The problem is that people are suggesting dye sub for quality and laser for
} quick & dirty. I need QUICK & QUALITY......and cheap. Printing on a dye
} sublimation printer would be too costly. There are some enhancement boards
} that can be added to laser printers, and I have a friend who has printed
} some of my scanned negatives on his Photoscan system, Philips's image
} recording system. They come close to answering the problem, but I am
} wondering what else there is out there. I believe that Philips's system may
} be LaserPix or PhotoJet Plus from XLI Corp.
}
} I realize that I may be asking you to repeat a conversation already carried
} out, but if you could pass on the consensus opinions, I would appreciate it.
}
}
} Joiner Cartwright, Jr., Ph.D.
}
} Director, Electron Microscopy tel.: (713)798-4658
} Department of Pathology, Rm.286-A FAX: (713)798-3945
} Baylor College of Medicine Internet: joiner-at-bcm.tmc.edu
} One Baylor Plaza Compuserve: 71555,1206
} Houston, Texas 77030 U.S.A.
}






From: James Patrick :      JPATRICK-at-OPUS.MCO.EDU
Date: Tue, 02 Apr 1996 13:50:44 -0500 (EST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: James Patrick :      JPATRICK-at-OPUS.MCO.EDU
Date: Tue, 02 Apr 1996 13:50:44 -0500 (EST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: Joiner Cartwright, Jr., Ph.D. :      joiner-at-bcm.tmc.edu
Date: Tue, 02 Apr 1996 12:46:48 -0600
Subject: Re: Printers for digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604021848.MAA01689-at-watson.bcm.tmc.edu}
X-Sender: joiner-at-bcm.tmc.edu
X-Mailer: Windows Eudora Version 2.1.1
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

At 08:11 AM 4/2/96 -0400, you wrote:

} Check out DOUBLE RES IV from Laser Printer Accessories Corp., a board you
} install into your current laserjet II or III to get 600 dpi quality. We have
} one and it really does make a BIG difference. 1-800-225-4098.
}
} Walter F. Bobrowski
} Subcellular Pathology
} Parke-Davis Pharmaceutical Research
} Ann Arbor, MI 48105
}
} TEL: 313-996-7814
} FAX: 313-996-5001
} E-Mail: BOBROWW-at-AA.WL.COM

****************************
Walter -

Is the 600 dpi output of this system better than the 600 dpi output of the
LaserJet 4 that we already have? There are boards that hot-rod the LaserJet
4 up to "2,400 equivalent dpi". One wonders if they will show "equivalent"
deposits in my kidney specimens. Thank you for your reply.


* * Joiner Cartwright, Jr. * *





From: Joiner Cartwright, Jr., Ph.D. :      joiner-at-bcm.tmc.edu
Date: Tue, 02 Apr 1996 12:46:38 -0600
Subject: Re: Printers for digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604021847.MAA01642-at-watson.bcm.tmc.edu}
X-Sender: joiner-at-bcm.tmc.edu
X-Mailer: Windows Eudora Version 2.1.1
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Scott -
Here is a reply that may interest you.
Joiner


At 09:35 AM 4/2/96 -0500, you wrote:
} In response to your recent message posted on the newsgroup concerning
} digital printers, there are several options. The first going to laser
} printers, and the second being dye sub printers. The cost of laser is
} obviously cheap, but not high enough quality perhaps for publication or
} reports where a photo is required. Dye sub printers ex: sony makes two
} digital dye sub printers. The first being model UP-D1500CN, it ouputs a
} picture size of approx. 3 5/8" x 4 3/4" which is near photographic quality.
} This printer sells for approx. $ 1,850.00 US and the print cost is approx.
} .86 cents. for color images. The other printer from sony is the UPD-8800
} with Interface card is approx. $7,300 US and the print cost for 8.5" x 11"
} is approx. $ 2.20 These printers will ouput a print in approx. 60 seconds.
} I do have sample prints if you require to see some.
}
} Sony makes a digital thermal printer that outputs approx. a polaroid size
} image on thermal paper in about 3.5 seconds, cost per print is approx. 9 cents.
}
} Epson makes some low cost printers, with a decent low cost image, the only
} problem with laser or ink jet with these images is that it takes quite a
} while to print.
}
} I don't know if I was helpful or not, but if I can be of any assistance,
} please feel free to contact me at your convenience.
}
}
} Best Regards,
}
} Philip Slakmon
} Scott Scientific
}
}
} _______________________________________________
} SCOTT SCIENTIFIC
} P.O. Box 66552 Station Cavendish,
} Montreal, Quebec, H4W 3J6, Canada
}
} Telephone: 514-485-2309 Fax: 514-485-9931
} Voice Mail: 514-888-6509
}
} WWW Site: http://www.ScottScientific.com
}
} E-Mail: slakmon-at-scottscientific.com
} info-at-scottscientific.com
} sales-at-scottscientific.com
} admin-at-scottscientific.com
} _________________________________________________
}
}
**********************************
Philip -

Thank you for your reply. Dyesub, as you see, may be capable of turning out
images of good enough quality to see, for example, the subtle
ultrastructural changes in kidney disease, but not at a price that we can
afford. And laser, although fast and cheap, cannot produce the image quality
necessary to show these subtle changes. One or more of the laser enhancement
boards MAY be the answer, but I am not yet convinced.


* * Joiner Cartwright, Jr. * *





From: Len Adleman :      adleman-at-pollux.usc.edu
Date: Tue, 2 Apr 1996 14:24:45 -0800 (PST)
Subject: Re: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: Joiner Cartwright, Jr., Ph.D. :      joiner-at-bcm.tmc.edu
Date: Tue, 02 Apr 1996 17:20:08 -0600
Subject: Re: Printers for digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604022321.RAA06357-at-watson.bcm.tmc.edu}
X-Sender: joiner-at-bcm.tmc.edu
X-Mailer: Windows Eudora Version 2.1.1
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

At 01:51 PM 4/2/96 EST, you wrote:

} I run a similar facility in the Pathology Department of a Veterinary
} College, but we also serve the entire College as well as some other
} parts of the university.
} With our SEM, we are entirely digital now...

*****************
Dr. Steffens -

Thank you for your reply. The savings are one of the main reasons that I
want to go digital. That includes savings in time as well as money. I am
considering the LaserPix. However there are some other technologies out
there that I want to look at as well.


* * Joiner Cartwright, Jr. * *





From: Jeff Allbright :      jeffallb-at-gol.com
Date: Wed, 3 Apr 1996 07:16:02 +-900
Subject: Kevex 8000 Data to PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {01BB212D.85B076C0-at-ppp236.gol.com}

Kevex still offers the 8000 to PC communication package using Kermit.
There is also a high speed data transfer option available using a SCSI
interface card. This is especially useful for transferring large image
files, etc. More information is available from Kevex Customer Service at
+1 (415) 591-3600.

- Jeff Allbright, Tokyo



Michael Lamvik asked about moving data from a Kevex 8000 to a PC. Kevex
used to sell a package to do this that cost about $1000. This bought you
a cable that connected to one of your printer ports on the 8000 and your
PC, a diskette with Kermit, software for the 8000, and a very thick manual.

I used it many times to move images from our 8000 to a Macintosh, and it
worked fine although it's not fast.

I don't know if Kevex sells this anymore since the 8000 is pretty old. A
gentleman named Robert Schaller was very helpful in getting us set up
with it, but this was at least 5 years ago and he may or may not still be
with Kevex.

Jim Stets
Air Products and Chemicals, Inc.
Allentown, PA
stetsjr-at-ttown.apci.com
My opinions are my own, not my employer's.








From: Jeff Allbright :      jeffallb-at-gol.com
Date: Wed, 3 Apr 1996 07:20:33 +-900
Subject: Kevex 8000 Data to PC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {01BB212E.209B9520-at-ppp239.gol.com}


Kevex still offers the 8000 to PC communication package using Kermit.
There is also a high speed data transfer option available using a SCSI
interface card. This is especially useful for transferring large image
files, etc. More information is available from Kevex Customer Service at
+1 (415) 591-3600.

- Jeff Allbright, Tokyo



Michael Lamvik asked about moving data from a Kevex 8000 to a PC. Kevex
used to sell a package to do this that cost about $1000. This bought you
a cable that connected to one of your printer ports on the 8000 and your
PC, a diskette with Kermit, software for the 8000, and a very thick manual.

I used it many times to move images from our 8000 to a Macintosh, and it
worked fine although it's not fast.

I don't know if Kevex sells this anymore since the 8000 is pretty old. A
gentleman named Robert Schaller was very helpful in getting us set up
with it, but this was at least 5 years ago and he may or may not still be
with Kevex.

Jim Stets
Air Products and Chemicals, Inc.
Allentown, PA
stetsjr-at-ttown.apci.com
My opinions are my own, not my employer's.









From: Buddy Steffens :      STEFFENS.B-at-calc.vet.uga.edu
Date: Tue, 2 Apr 1996 13:51:24 EST
Subject: Re: Printers for digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

microscopy-at-Sparc5.Microscopy.Com

} I run an EM lab in a Pathology department of a medical school. Most of my EM
} is diagnostic work on kidney specimens and tumors. We printed out 4,500
} images for each of the two past years. I would like to go digital, but
} finding a printer that will handle this load cheaply and quickly, yet with
} the required high quality is difficult. We plan to produce the usual 3 1/4"
} X 4" negatives in our scope as we always have. Then we would scan the
} negatives into a computer (IBM clone), storing them temporarily on a large
} hard disk, and ultimately archiving them on a writable CD system. The prints
} need not be durable as once they are studied and the case closed, they would
} be trashed. The negatives would be archived as would be the image files on
} the CD's.
}
} The problem is that people are suggesting dye sub for quality and laser for
} quick & dirty. I need QUICK & QUALITY......and cheap. Printing on a dye
} sublimation printer would be too costly. There are some enhancement boards
} that can be added to laser printers, and I have a friend who has printed
} some of my scanned negatives on his Photoscan system, Philips's image
} recording system. They come close to answering the problem, but I am
} wondering what else there is out there. I believe that Philips's system may
} be LaserPix or PhotoJet Plus from XLI Corp.
}
I run a similar facility in the Pathology Department of a Veterinary
College, but we also serve the entire College as well as some other
parts of the university.
With our SEM, we are entirely digital now...I don't even support film
on it. With the TEM, we have a side-mounted digital camera system
(JEOL TV-CAM) that is used for image acquisition, but we still use film for publication
purposes. Most of the TEM images are for study purposes only, and
for this we dump the image (using the system's software) to a
Laserjet 4 printer through a Laserpix board. After a bit of
processing, ie BCG adjustment, unsharp mask sharpening, etc., you get
a "work print" that is usually acceptable. For higher quality, we
also have a Codonics dye sublimation printer, on which you can put 4
images per page. For TEM, its certainly better quality than the
Laserjet, but is usually not publication quality unless you do some
intensive processing. SEM and light micrographs on it however are
certainly publication quality.
Since most images we take will not be published, we get along fine
with the Laserjet prints for TEM. We do use a high quality glossy
paper in it which drives up the cost of the prints to about .12 each.
That still beats the 2.00+ of the Codonics.
Since going mostly digital, our TEM film images have dropped from
about 2000 per year to about 200 per year. The savings in tech
labor, direct costs, and chemical wastes are quite substantial.




W. L. Steffens, Ph.D
Dept. of Veterinary Pathology
College of Veterinary Medicine
University of Georgia
Athens, GA 30602
STEFFENS.B-at-CALC.VET.UGA.EDU
Voice: (706) 542-5536
FAX: (706) 542=5828




From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Tue, 2 Apr 1996 19:18:16 -0600
Subject: Re: EM: Marine invertebrate texts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} } Dear Microscopists,
} } Does anyone out there know of a textbook or any other reference material
} } which covers integumental ultrastructure - cuticle, epithelium and
} } associated structures and also oogenesis - in marine invertebrates. We hav=
e
} } a marine science student who is studying subclass Ascothoracida who is
} } having problems getting the ultrastructure info she needs.
}
} You might browse the "Microscopic Anatomy of Invertebrates" series
} (Frederick W. Harrison, Burton J. Bogitsh, Eds.), Wiley-Liss, New York,
} 1991. It=B4s a highly comprehensive series of 15 volumes, covering in-depth
} ultrastructure from Protozoa up to Hemichordata, Chaetognatha, and else.
}
} -Dietmar-

As well as this ref, there is:

Neville, A.C. 1975. _The Biology of the Arthropod Cuticle_. Vol,
4/5 in the series "Zoophysiology and Ecology". Springer-Verlag.

Nothing on Ascothoracida in specific, but a good general ref, and
literature review up to about 1974.
Phil

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: Joiner Cartwright, Jr., Ph.D. :      joiner-at-bcm.tmc.edu
Date: Tue, 02 Apr 1996 17:20:17 -0600
Subject: Printers for digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604022321.RAA06384-at-watson.bcm.tmc.edu}
X-Sender: joiner-at-bcm.tmc.edu
X-Mailer: Windows Eudora Version 2.1.1
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I recently posted a query concerning printers suitable for producing high
quality monochrome images from micrographs scanned into a computer....a
printer that would allow me to "go entirely digital" and get me out of the
darkroom, at least for all imaging except the most demanding applications. I
have received a number of replies. Boiled down, the answers include:

1.) For quick & dirty, but usable work prints, a laser printer producing
600+ dpi;

2.) For high quality, a dye sublimation printer; and

3.) An intermediate would be a laser printer with one of the after market
enhancement boards that runs the lasers resolution up to 1,200 dpi or "2,400
dpi equivalent", whatever that means;

4.) A fourth suggestion two or three people made was the Harris PhotoPro
2000 Gray Scale Digital Printer. This is a $10,000.00 silver salt laser
printer that claims only 256 dpi, but each dot has 256 gray levels
available. It was stated that the cost/print was less than $1.00.

Do any of you have any experience with this latter machine, and would you be
willing to share your thoughts? Does the increased contrast resolution make
up for the POSSIBLY low spatial resolution? Do I really want to spend ten
big ones on a black & white printer?




Joiner Cartwright, Jr., Ph.D.

Director, Electron Microscopy tel.: (713)798-4658
Department of Pathology, Rm.286-A FAX: (713)798-3945
Baylor College of Medicine Internet: joiner-at-bcm.tmc.edu
One Baylor Plaza Compuserve: 71555,1206
Houston, Texas 77030 U.S.A.






From: stpiera-at-lepton.hmco.com (Aaron Stpierre)
Date: Tue, 2 Apr 1996 22:48:15 -0500 (EST)
Subject: unsub

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: Peters-at-BSAC.UCHC.EDU (Klaus-Ruediger Peters)
Date: Tue, 2 Apr 1996 20:53:19 -0500
Subject: Re: Printers for digital images: Tests

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d01ad87835f46cf-at-[155.37.2.10]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

In response to last years questions raised on the performance of digital
printers I tested several printers with a 1Kx1K image. Printer performances
were compared at the level of whole page images and 10x enlargements of the
printed images (LaserJet, Lazar Print, Dye sub,). Results are available at

http://panda.uchc.edu/htklaus/DigiLab/Printing.html

In conclusion, you have to ask how many lines per inch are printed and at
how many gray levels (not always the same as dots per inch), plain paper or
special paper, permanence of the print, printing time, printing costs. I
print 300 lpi at 256 gray levels (4,800 dpi) on plain paper in 20-45 sec
for 1-10 MB of images per page at 3 cents per page with a HP LaserJet and
enhancer board in archival quality.

Best regards Klaus


*****************************************

} At 08:11 AM 4/2/96 -0400, you wrote:
}
} } Check out DOUBLE RES IV from Laser Printer Accessories Corp., a board you
} } install into your current laserjet II or III to get 600 dpi quality. We have
} } one and it really does make a BIG difference. 1-800-225-4098.
} }
} } Walter F. Bobrowski
} } Subcellular Pathology
} } Parke-Davis Pharmaceutical Research
} } Ann Arbor, MI 48105
} }
} } TEL: 313-996-7814
} } FAX: 313-996-5001
} } E-Mail: BOBROWW-at-AA.WL.COM
}
} ****************************
} Walter -
}
} Is the 600 dpi output of this system better than the 600 dpi output of the
} LaserJet 4 that we already have? There are boards that hot-rod the LaserJet
} 4 up to "2,400 equivalent dpi". One wonders if they will show "equivalent"
} deposits in my kidney specimens. Thank you for your reply.
}
}
} * * Joiner Cartwright, Jr. * *

******************************************************************************
* : *
* Klaus-Ruediger Peters, Ph.D. : WWW Home Page: *
* Director, Molecular Imaging Laboratgory : *
* Biomolecular Structure Analysis Center : Molecular Imaging Laboratory *
* University of Connecticut Health Center : http://panda.uchc.edu/ *
* 263 Farmington Ave. : htklaus/index.html *
* Farmington, CT 06030-2017; U.S.A : Differential Hysteresis *
* : Processing Demo at http:// *
* Tel: (203) 679-3977; Fax: (203) 679-1989 : panda.uchc.edu./htbit/indiv/ *
* e-mail: Peters-at-BSAC.UCHC.EDU : /software_docs/dhp.html *
* : *
****************************************************************************
**






From: ota-at-sun3.oulu.fi (Olli Taikina-aho)
Date: Wed, 3 Apr 1996 09:42:54 +0300 (EET DST)
Subject: point counter needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello friends

Could someone tell me if there is any company delivering mechanical point
counter systems. I need it for a polarizing microscope for geological samples.
Olli Taikina-aho
University of Oulu
Institute of Electron Optics
Box 400
SF-90571 Oulu
ota-at-cc.oulu.fi
tel. +358-81-5533142
fax. +358-81-5533149





From: houpt-at-worldaccess.nl (houpt)
Date: Wed, 3 Apr 1996 12:24:27 +0100
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604031024.KAA25184-at-worldxs.worldaccess.nl}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

unsubscribe

BIOMET
P.M.Houpt
The Hague.
The Netherlands.






From: French Michele H :      French_Michele_H.PRILVMS3-at-msmail.bms.com
Date: Wed, 03 Apr 1996 09:16:24 +0800 (U)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Suscribe french_michele-h.PRILVMS3-at-MSMAIL.BMS.COM




From: Jose'Angel Conchello :      josec-at-abbe.wustl.edu
Date: Wed, 3 Apr 1996 10:41:40 -0600
Subject: BiOS '97 Call For Papers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Three-Dimensional Microscopy: Imaging Acquisition and Processing IV

Conference Chairs: Carol J. Cogswell, Univ. of Sydney (Australia); Jose-Angel
Conchello, Washington Univ.; Tony Wilson, Univ. of Oxford (UK)

Program Committee: Alan Bearden, Univ. of California/Berkeley; G. J.
Brakenhoff, Univ. of Amsterdam (Netherlands); Kjell Carlsson, Royal Institute
of Technology (Sweden); ; Seth Goldstein, National Institutes of Health;
Gordon S. Kino, Stanford Univ.; Andres Kriete, Univ. Giessen (FRG); Frederick
Lanni, Carnegie Mellon Univ.

This conference will explore the rapidly developing field of three-dimensional
microscopy. Consideration will be given to the characteristics of the overall
system design, as well as to the topics of image formation, image recording,
deconvolution in 2,3 or more dimensions, and digital methods of producing and
displaying the resulting 3D reconstruction. Recent innovations in 3D
microscopy are having a large impact in the biological and medical fields, as
well as in materials science and the semiconductor industry. Many microscopes
are now fully integrated systems, including computer hardware and software. It
is hoped that the broad range of relevant topics being presented at this
symposium will serve to encourage interaction among instrumentation engineers,
computer image analysts, and researchers in the various fields of application.

Papers are invited in the following and related areas:
- confocal microscopy
- 3D image formation
- instrumentation for 3D microscopy
- time-resolved image acquisition systems
- image processing and analysis
- 3D image reconstruction
- deconvolution in 2/3 or more dimensions
- super-resolution
- spatiotemporal reconstruction of living cells and tissues
- applications of 3D microscopy in materials science
- profilometry
- image visualization techniques for 3D microscopy systems, including volume
rendering, animation, stereoscopic and holographic displays



This conference is just one of 27 topics to be held at the BiOS '97 symposium,
which is part of the larger event called Photonics West '97:
8-14 February 1997
San Jose Convention Center
San Jose, California USA

TO OBTAIN ALL CALLS FOR PAPERS ELECTRONICALLY
The calls for papers for all conferences in the BiOS '97 symposium (as well as
all other symposia that are part of Photonics West) will be available
1 May on the SPIE Web site
http://www.spie.org/web/meetings/calls/submissions.html;
by anonymous FTP
ftp://spie.org/meetings/calls/PW97_readme.txt;
or by e-mail file retrieval
send a message to info-spie-request-at-spie.org with the following in the message
body:
send [meetings.calls]PW97_readme.txt

For a printed BiOS '97 call for papers or other information:
E-mail: PW97-at-spie.org
Fax: 360/647-1445
Phone: 360/676-3290

IMPORTANT DEADLINES
Paper Abstracts Due from Authors:
15 July 1996 (post-meeting proceedings)
Advance Programs due from Chairs:
12 August 1996 (post-meeting proceedings)
Camera-Ready Abstracts due from Authors:
9 December 1996
Manuscripts Due from Authors:
13 January 1997 (post-meeting proceedings)

GUIDELINES FOR SUBMITTING AN ABSTRACT

Send a 250 word abstract of your paper, by 15 July 1996, in ONE of the
following ways:

* SPIE WEB - Complete the convenient form found at SPIE Web site:
http://www.spie.org/web/meetings/calls/submissions.html

* or E-MAIL each abstract separately to: abstracts-at-spie.org in ASCII text
(not encoded) format. To ensure receipt and proper processing of your
abstract, write on SUBJECT line: PW '97 and Conference Chair.

Example: SUBJECT: PW97 (John Smith)

* or MAIL three copies of your abstract to:
BiOS '97
SPIE, P.O. Box 10, Bellingham, WA 98227-0010 USA
Shipping address: 1000 20th St., Bellingham, WA 98225 USA
Telephone 360/676-3290

* or FAX one copy to SPIE at 360/647-1445 (send each abstract separately).


Be sure each abstract includes the following:

1. CONFERENCE CHAIR and CONFERENCE TITLE (submit to ONLY ONE conference)
to which the abstract is submitted

2. AUTHOR LISTING (List principal author first)
for each author: full name and affiliation, mailing address, phone/fax
numbers, email

3. ABSTRACT/PAPER TITLE

4. ABSTRACT TEXT: 250 words

5. KEYWORDS: maximum of 5 keywords

6. BRIEF BIOGRAPHY of the principal author: 50-100 words

Please contact SPIE if you have any questions or require further information.


Marilyn E. Gorsuch, Technical Programs Manager
SPIE
PO Box 10
1000-20th Street
Bellingham, WA 98225
Ph: 360/676-3290
Fax: 360/647-1445
e-mail: marilyn-at-spie.org





From: David.Rothbard-at-ipst.edu (David Rothbard)
Date: Wed, 3 Apr 1996 12:47:28 -0500
Subject: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v0213050bad8867979a4c-at-[199.77.235.102]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Like many of you , I have worked in EM laboratories with water (fire
activiated) sprinklers overhead. In some cases I have successfully lobbied
to have them removed, as I felt there was a greater danger of electrical
shock or massive instrument damage if they tripped in error (beware the
pipefitter). However, there is always the possibility that your EM could
be the source of a fire that causes great damage and injury. How do people
feel about this?

David Rothbard

--
Institute of Paper Science and Technology






From: samso-at-orkney.ph.albany.edu
Date: Wed, 03 Apr 1996 13:28:24 EST
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: John M. Libert :      jlibert-at-cpcug.org
Date: Wed, 03 Apr 1996 11:53:52 -0800
Subject: Re: point counter needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {3162D750.4D11-at-cpcug.org}

Olli Taikina-aho wrote:
}
} Hello friends
}
} Could someone tell me if there is any company delivering mechanical point
} counter systems. I need it for a polarizing microscope for geological samples.
} Olli Taikina-aho
} University of Oulu
} Institute of Electron Optics
} Box 400
} SF-90571 Oulu
} ota-at-cc.oulu.fi
} tel. +358-81-5533142
} fax. +358-81-5533149Yes. Prior Scientific, Inc of Rockland, MA sells an electro-mechanical
pointcounter. It has a number of nice features and is fairly simple. Give
them a call at 617-878-8442 (FAX 617-878-8736).

John M. Libert
OPELCO
OPtical ELements COrporation
Sterling, VA.
703-471-0080 Ext. 217




From: Leslie Gartner :      lgartner-at-umabnet.ab.umd.edu
Date: Wed, 3 Apr 1996 15:24:03 -0500 (EST)
Subject: HELP Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I apologize for posting this to the entire community, but I no longer
have the majordomo address to unsubscribe nor the proper procedure.
Please e-mail me directly to reduce bandwidth clutter.

Thank you,

Les Gartner
lgartner-at-umabnet.ab.umd.edu







From: David.Rothbard-at-ipst.edu (David Rothbard)
Date: Wed, 3 Apr 1996 16:00:50 -0500
Subject: Re: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v0213050cad8896f9b99f-at-[199.77.235.102]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Several people have mentioned Halon to me, but I was under the impression
that Halon has fallen out of favor due to potential toxicity and
suffocation.

David Rothbard

--
Institute of Paper Science and Technology






From: rnessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Wed, 3 Apr 1996 15:51:49 -0600 (CST)
Subject: Re: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Wed, 3 Apr 1996, David Rothbard wrote:

} Like many of you , I have worked in EM laboratories with water (fire
} activiated) sprinklers overhead. In some cases I have successfully lobbied
} to have them removed, as I felt there was a greater danger of electrical
} shock or massive instrument damage if they tripped in error (beware the
} pipefitter). However, there is always the possibility that your EM could
} be the source of a fire that causes great damage and injury. How do people
} feel about this?
}
} David Rothbard
}
} --
} Institute of Paper Science and Technology
}
Our lab is equipped with a Halon fire suppression system. I can't
verify that it works, as we have never needed it :)


Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu





From: Joyce Craig :      bafpjec-at-uxa.ecn.bgu.edu
Date: Wed, 3 Apr 1996 12:04:28 -0600 (CST)
Subject: LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are using LR white for the first time in anticipation of doing some
immunocytochemistry and also because I have several people who are
working with bacteria and fungi.
We have been fixing the bacteria in 2 + 2, trying different
concentrations of buffer because the bacteria grows in a 25% sodium
chloride. Then we mix the bacteria with agar, fix again, osmicate, and
dehydrate to 70%. Then we mix the LR White with 70 % for infiltration, then
straight LR White finally flat embed in LR White and polymerize at 60
degreesC.
We wrap the embedding molds in Saran wrap (not an off-brand) before
polymerization.
The blocks in blue JB4 molds hardened, but have a surface stickiness.
The blocks in the clear molds disappeared. The black blocks of osmicated
agar are there but the epoxy is gone.
Does anyone have any thoughts on this?
I really like running up nice blocks of pancreas and kidney and even
cardiac muscle in epoxy and taking pretty pictures in the microscope but
unfortunately no-one here is doing this nice rewarding basic work.




From: vit-at-felix.scvnet.com (Jim Yankovich)
Date: Wed, 3 Apr 1996 17:31:37 -0800
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please Unsubscribe






From: rgwhite-at-vaxc.cc.monash.edu.au (Rosemary White)
Date: Thu, 04 Apr 1996 09:16:15 +1200
Subject: Postdoc - TEM, fluorescence (maybe confocal)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


POSTDOCTORAL FELLOWSHIP IN PLANT CELL BIOLOGY

Applications are invited for a postdoctoral fellowship on a 3-year project
investigating computer modelling of macromolecular structures in plant
cells. Part of this work requires preparation of TEM and
immunofluorescence micrographs for comparison with computer generated
images of plant cytoskeleton and plasmodesmata. We are also investigating
various aspects of the interactions between these cell components and their
dynamic behaviour in cells. Some confocal fluorescence microscopy may be
required.

Applicants should have experience in TEM and immunofluorescence of plant
material. Experience in freeze-substitution techniques and
immunofluorescence of the plant cytoskeleton would be an advantage.

The position is available from 1 July, and the successful candidate will be
expected to take up the appointment by 1 September 1996 at the latest.
Salary will commence at $(Australian)37,345. The appointment will be for
12 months in the first instance, renewable for a further 12 months (longer
if new grant application is successful). Some remuneration for removal
expenses will be available.

Applications should arrive by 4 May 1996, with CV naming 2 referees.

For further information, contact


Rosemary White
__ /
Department of Ecology _/ \__/ \
and Evolutionary Biology / \
Monash University / Australia \
Clayton, Victoria 3168 \ ____ /
phone 61-3-9905 5670 \_/ \_*_/
fax 61-3-9905 5613 __
email rgwhite-at-vaxc.cc.monash.edu.au \/






From: Sarka Lhotak :      lhotaks-at-fhs.csu.mcmaster.ca
Date: Wed, 3 Apr 1996 21:21:55 -0500 (EST)
Subject: immuno:source of antibodies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear immunobuffs,

I was very impressed with the ChemExpo www site that
was recommended here recently. I wonder whether there is
something similar from antibody producers that would enable to
locate a source of a particular antibody easily. My search for
this was unsuccessful. Thanks for your help.

Sarka Lhotak
EM Facility, McMaster University
Hamilton, Ontario, Canada
lhotaks-at-fhs.mcmaster.ca




From: Bob Citron :      Bob_Citron-at-cc.chiron.com
Date: 4/3/96 3:48 PM
Subject: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Like many of you , I have worked in EM laboratories with water (fire
activiated) sprinklers overhead. In some cases I have successfully lobbied
to have them removed, as I felt there was a greater danger of electrical
shock or massive instrument damage if they tripped in error (beware the
pipefitter). However, there is always the possibility that your EM could be
the source of a fire that causes great damage and injury. How do people
feel about this?

David Rothbard

--
Institute of Paper Science and Technology







From: Microscopy-request
Date: Wednesday, April 03, 1996 5:16PM
Subject: Re: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


David;

A few years ago, we had a very bad rainstorm in Southern California. The
roof
of our R&D building (which houses the SEM) leaked, and the ceiling tile got
so
soaked that it fell in on the SEM console. When I came in the next morning,

there was a steady stream of water pouring into the keyboard and down into
the
electronics of the SEM. Obviously, the SEM had tripped off. I unplugged it
and
called the vendor (Cambridge, at the time), since we were under a service
contract. End result - we allowed it to dry for 3 days and fired it up. It

worked fine, and has been operational ever since; no problems have been
attributed to this incident. In summary, I believe that the risks
associated
with fire far outweigh the risks associated with water.

Regards,

Bob




From: Schoonhoven, Robert :      rschoonh.sph-at-mhs.unc.edu
Date: Thu, 4 Apr 1996 8:07:25 -0500
Subject: Re: immuno:source of antibodies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {85C4633101F70300-at-mhs.unc.edu}
In-Reply-To: {7DC4633101F70300}

I haven't found anything on the net but an excellent (and not to
expensive ) refferance is "Linscotts Directory of Immunological and
Biological Reagents" the send out constant updates and is available in
written and PC compatible form. The phone # for additional information
is 707-544-9555.

regards, bob

Robert Schoonhoven
Laboratory of Molecular Carcinogenesis and Mutagenesis
University of North Carolina
CB#7400, Rosenau Hall
Chapel Hill, NC 27599-7400
Ph. 919-966-6343
Fax 919-966-6123





From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Thu, 4 Apr 1996 08:20:06 GMT
Subject: Re: LR White

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I maintain an archive of most of the biological EM related
material that comes across this list. Since you have net acess I will assume
you can get acess to the www. Go to the web page listed at the bottom of
this message and click of the "Tips & Tricks Wizard" In there is a file
called "Embedding with LR White" which may be of use to you. If for some
reason you do not have acess to the web or cannot read the file, let me know
and I will e-mail it to you .



At 12:04 PM 4/3/96 -0600, you wrote:
} We are using LR white for the first time in anticipation of doing some
} immunocytochemistry and also because I have several people who are
} working with bacteria and fungi.
} We have been fixing the bacteria in 2 + 2, trying different
} concentrations of buffer because the bacteria grows in a 25% sodium
} chloride. Then we mix the bacteria with agar, fix again, osmicate, and
} dehydrate to 70%. Then we mix the LR White with 70 % for infiltration, then
} straight LR White finally flat embed in LR White and polymerize at 60
} degreesC.
} We wrap the embedding molds in Saran wrap (not an off-brand) before
} polymerization.
} The blocks in blue JB4 molds hardened, but have a surface stickiness.
} The blocks in the clear molds disappeared. The black blocks of osmicated
} agar are there but the epoxy is gone.
} Does anyone have any thoughts on this?
} I really like running up nice blocks of pancreas and kidney and even
} cardiac muscle in epoxy and taking pretty pictures in the microscope but
} unfortunately no-one here is doing this nice rewarding basic work.
}
}







} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {

Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 904-392-1295
ICBR EM Core Lab fax 904-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/





From: Giorgio :      giocar-at-risc990.bologna.enea.it
Date: Thu, 4 Apr 1996 15:12:50 GMT
Subject: answers to: need books on biology...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Following my request for more details, Ms Kaye Patterson
{kayep-at-deakin.edu.au} has sent me more information about a book I think can
be very interesting for many people. For its possible general interest, I
send this message to the list.
Greetings George
----------------------------------------------------------
George,
...
A company in America called Ward's Biology has a very large
catalogue which they send to us here in Australia every year. Their address is
P.O. Box 92912
Rochester, NY 14692-9012.
For International Customers:
Fax: 716-334-6174
Phone 716-359-2502.
They have a large selection of Biological books, manuals, guides,
etc. which you would probably find useful, as well as models, equipment,
charts, microscope slides etc. etc. etc.
They have the book you are asking about available for US$65.00,
Catalogue No. 32 W 0902.
The full details of the book are:
Morholt, Evelyn and Brandwein, Paul F. (1986) A Sourcebook for the
Biological Sciences. 3rd ed.
The publisher of my 2nd ed. copy is Harcourt, Brace & World, Inc.
New York, but it was published in 1966 (and also has Alexander Joseph as a
third author), so I don't know what the publishing details for the 3rd ed. are.
The Ward's catalogue says of the book "No biology teacher should be
without this comprehensive sourcebook of techniques, procedures,
demonstrations, projects, experiments and support material. Also includes
sections on maintaining organisms, safety and chemical preparations.
Hardcover, 813 pages, with illustrations."
I can only comment about the book as a biology technician and not as
a teacher. I find it useful as the book details activities and experiments
which our Uni students do, along with the practical aspects which I need to
get the classes ready. eg. for classes looking at Protozoans,
invertebrates, etc., the book describes the best way to get material from
the field, how to culture the various groups, how to prepare material for
microscopic examination, anesthetization, staining and fixation techniques.
The book covers many areas of Biology - anatomy, cells & tissues,
photosynthesis, ingestion & digestion, transpiration, circulation,
respiration, behaviour & coordination, genetics, growth & differentiation,
evolution, and ecology. There is a section on special techniques -
Maintaining animals useful in the classroom, Growing plants useful in the
classroom and Stockroom and facilities for biology. This is the section I
use the most.
Hope you find all this useful.
Kaye





From: dbd1-at-uclink4.berkeley.edu
Date: Thu, 4 Apr 1996 09:30:43 -0800 (PST)
Subject: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Halon is great for stopping fires but yes, the same mechanism that
extinguishes the fire will suffocate people. Sprinklers are generally OK
but I prefer a standard A-B-C rated fire extinguisher in the scope room.
We have had "umbrellas" mounted over the scopes for the last 20 years-
aluminum tube framework covered by a heavy-duty waterproof nylon fabric.
We reside in an old building and very occasionally, plumbing accidents
result in water leaking throught the ceiling above the scope.
Regards,


Doug Davis
Staff Research Associate
Electron Microscope Facility
University of California
Berkeley, CA 94720
(510) 642-2085, fax 643-6207
dbd1-at-uclink4.berkeley.edu







From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Thu, 4 Apr 1996 12:04:35 -0600
Subject: EM and Water

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130500ad89bd67f2c4-at-[131.230.97.69]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I would venture that analog instruments might be able to weather the
onslaught of water better than newer instruments with lots of sensitive
electronics. The best approach would be to contact computing centers and
see what measures are taken to protect their instrumentation.

Secondly, some sort of risk assessment needs to be considered here. What
are the odds that floods will be generated needlessly versus having a
facility or instrumentation destroyed by fire? Most instances of fires
affecting EM units that I have heard about were caused by fires originating
other than in the EM room. Perhaps sprinklers could be used in locations
other than the EM while the EM rooms could be protected by more
electronically-friendly agents (Halon, carbon dioxide, etc).

Just my thoughts .....


#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Thu, 4 Apr 1996 12:04:35 -0600
Subject: EM and Water

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130500ad89bd67f2c4-at-[131.230.97.69]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I would venture that analog instruments might be able to weather the
onslaught of water better than newer instruments with lots of sensitive
electronics. The best approach would be to contact computing centers and
see what measures are taken to protect their instrumentation.

Secondly, some sort of risk assessment needs to be considered here. What
are the odds that floods will be generated needlessly versus having a
facility or instrumentation destroyed by fire? Most instances of fires
affecting EM units that I have heard about were caused by fires originating
other than in the EM room. Perhaps sprinklers could be used in locations
other than the EM while the EM rooms could be protected by more
electronically-friendly agents (Halon, carbon dioxide, etc).

Just my thoughts .....


#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: Weiming Yu :      weiming-at-LFD.physics.uiuc.edu
Date: Thu, 4 Apr 1996 12:52:38 -0600 (CST)
Subject: Re: Temperature Controller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi Friends,

Thank your all for your responses. That is a great help.


Weiming Yu Ph.D.
Dept of Physics, UIUC
1110 W. Green Street
Urbana, IL 61801
w-yu-at-uiuc.edu




From: wise-at-vaxa.cis.uwosh.edu
Date: Thu, 04 Apr 1996 09:46:53 +0000
Subject: Re: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} David;
}
} A few years ago, we had a very bad rainstorm in Southern California. The roof
} of our R&D building (which houses the SEM) leaked, and the ceiling tile got so
} soaked that it fell in on the SEM console. When I came in the next morning,
} there was a steady stream of water pouring into the keyboard and down into the
} electronics of the SEM. Obviously, the SEM had tripped off. I unplugged it
} and
} called the vendor (Cambridge, at the time), since we were under a service
} contract. End result - we allowed it to dry for 3 days and fired it up. It
} worked fine, and has been operational ever since; no problems have been
} attributed to this incident. In summary, I believe that the risks associated
} with fire far outweigh the risks associated with water.


Yes, but rainwater is clean. I have been told (third- or fourth-hand story
by now) that it is not the water in sprinkler systems as much as the rust,
junk and (perhaps) rust inhibitors in the pipes that cause the major
damage.

Bob Wise






From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 4 Apr 1996 09:24:27 -0500 (EST)
Subject: Re: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Like many of you , I have worked in EM laboratories with water (fire
} activiated) sprinklers overhead. In some cases I have successfully lobbied
} to have them removed, as I felt there was a greater danger of electrical
} shock or massive instrument damage if they tripped in error (beware the
} pipefitter). However, there is always the possibility that your EM could
} be the source of a fire that causes great damage and injury. How do people
} feel about this?
}
Dear David,
We have had one fire in an EM room--started when the LN delivery
system ran dry and a solenoid valve shorted out. There are definitely
other potential fire dangers, such as vacuum pumps. We do not have
sprinklers. I have not heard of erronious tripping of sprinkeler sys-
tems being a particular problem. All-in-all, I lean toward having a
fire supression system, but there are obvious problems with a water-
based one, even if it trips due to a real fire. Is there a better
system for areas where large voltages and/or currents are present?
Perhaps some industries have solved this problem.
Yours,
Bill Tivol




From: Jay Jerome :      jjerome-at-bgsm.edu
Date: Thu, 4 Apr 1996 16:34:35 -0500 (EST)
Subject: Re: EM and Water

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




With regard to the thread about fire extinquishers in EM suites, we were
told during our last safety inspection that we would need to get rid of
our halon system the next time it needed a charge. This was mostly
because the insurance companies would no longer cover such things. They
require springler systems. They would rather pay the for lost
equipment than the hazards of halon. Of course our safety officer had no
answer to my question of whether the school would pay the 5,000
deductible or whether I would have to pay this.

Jay
Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-bgsm.edu *
**************************************************************





From: emccjb-at-ttuhsc.edu (Charles J. Butterick)
Date: Thu, 4 Apr 1996 08:54:30 -0600
Subject: LN2 Safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
I've read with interest the comments of others questioning liquid
nitrogen safety regulations. In spite of many years of pratical experience
with the stuff and being a trained safety officer, I took some time to
examine the subject. I disagree with those who have problems with the
regulations. The regs seem to be aimed at preventing continuous contact
and the resultant injuries.
In the CRC Handbook of Laboratory Safety, 3rd edition, pages
315-317, lists all the basic precautions (face shield, impervious clothing,
etc.). For those who disagree, gloves are optional.....except, as Tivol
pointed out, when in contact with cold metal parts. The CRC actually
suggests using some sort of potholder to protect hands from the metal. The
CRC also points out that efforts should be made to prevent trapping the
cryogen. When trapped against the skin, LN2 will cause injury since the
contact is no longer momentary and becomes continuous contact. Cuffless
pants worn over the shoes/boots are reccommended, i.e.spilled LN2 that
gets trapped inside a shoe with the foot may cause injury. For those who
suggest sandals as appropriate should think about the possibility that with
a major spill, some LN2 could be trapped beneath the foot or under the
strap of a sandal, possibly causing injury. Appropriate shoes can protect
the foot for short periods. If the footwear is frozen, generally it can be
removed before injury occurs. If gloves are used, the fit should be loose
so that they may be shed quickly...ala a hockey player's actions when he
has been treated less than courteously by an opposing player. The problem
I have with faculty and staff is the use of latex gloves while using LN2.
The Leidenfrost effect can definitely cause injury if the latex freezes
against the skin. For those who would not wear appropriate gloves, have
you considered the possibility of LN2 being trapped under a watch band, or
under a ring? In a major splash, could LN2 become trapped at the beltline,
between the pants and shirt? The whole idea is to make sure that most any
cryogen does not get trapped against the skin. If you can handle LN2
safely without protection, or have done it for years without incident, more
power to you. I hope your luck holds out.
Interestingly enough, Air Liquide, a national supplier of cryogens,
also lists similar precautions in the use of liquid nitrogen and other
cryogens.
Texas Tech has (arguably) a large number of beautiful women, but
they are not the ones coming to get liquid nitrogen....and of the
individuals who do get LN2 from our lab, I would rather them come clothed.



Charles J. Butterick (Chuck)
Electron Microscopy Center
Department of Cell Biology
and Biochemistry
Texas Tech University Health
Sciences Center
3601 4th Street
Lubbock, Texas 79430

vox (806) 743-1633
fax (806) 743-1219
email emccjb-at-ttuhsc.edu or
chuck-at-micron1.lubb.ttuhsc.edu






From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Thu, 4 Apr 1996 17:05:49 GMT
Subject: EMs and sprinklers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The sprinkler next to your scope may not be the problem. We are on
the second floor, but when the sprinklers went off on the sixth floor, water
eventually worked its way down to us. Luckily this happened in the daytime
and we had enough time to cover done the scopes and computers. No one knew
how to turn off the water so there was 6 inches on the 6th floor and
ceilings collapsed on the 5th floor exactly as they are designed to do
during a fire. There was no fire.
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-





From: BGIAMMARA-at-magnum.mco.edu
Date: Thu, 04 Apr 1996 11:56:26 -0500 (EST)
Subject: Sprinklers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


David,
We had a great central laboratory in the basement of the Graduate School
with a fire sprinkler above each of three electron microscopes. The Dean
worried that the sprinklers would ruin the microscopes, so they got
insurance on them and required that I keep large plastic tarps to cover
them in case the sprinklers went off. They never did, howwever...
One Saturday, we had a huge rainstorm and the Dean's secretary called
me saying their seemed to be a "little problem in the lab." I rushed over
there to find that the drains at the top of the stairs had filled with
leaves and likewise the drain at the bottom and huge amounts of water,
like a waterfall, was cascading down the steps and filling up the lab.
The building power had gone off and the emergency lighting was on in the
hallways. I called for the maintenance crew, but they were priortized
to go save "expensive" student computers in another building. So there
I was with my broom, frantically cleaning drains and pushing water, feeling
like a character in Walt Disney's Fantasia. Because of this long and
intense fight, the only thing we lost was the carpeting, papers and books
on the lowest shelf. Saved the floor tile. The minimal rust on the feet of
the EM's wasn't even noticeable. The lab was fine. Sprinklers a non-issue.
The real devastation was done by the administration, later, when a
new Dean said "What are electron microscopes good for?" eyeing the space and
salary line he could use for a new Executive Secretary. But that's another
story. Kind regards and good wishes to all, Beverly






From: BGIAMMARA-at-gemini.mco.edu
Date: Thu, 04 Apr 1996 11:56:26 -0500 (EST)
Subject: Sprinklers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



----- Begin message from IN%"postmaster-at-gemini.mco.edu" 4-Apr-96


Return-path: {}
Received: from gemini.mco.edu by gemini.mco.edu (PMDF V4.3-7 #8091)
id {01I355J900A88Y5QD6-at-gemini.mco.edu} ; Thu, 4 Apr 1996 11:56:37 EST


--Boundary (ID JlM655DZrjy0J4/C5dhX7g)
Content-type: TEXT/PLAIN; CHARSET=US-ASCII

The message could not be delivered to:

Addressee: David.rothbard-at-ipset.edu
Reason: Illegal host/domain name found.


--Boundary (ID JlM655DZrjy0J4/C5dhX7g)
MIME-version: 1.0
Content-type: MESSAGE/RFC822

Received: from gemini.mco.edu by gemini.mco.edu (PMDF V4.3-7 #8091)
id {01I355J6DYVA8Y5MGV-at-gemini.mco.edu} ; Thu, 4 Apr 1996 11:56:26 EST

David,
We had a great central laboratory in the basement of the Graduate School
with a fire sprinkler above each of three electron microscopes. The Dean
worried that the sprinklers would ruin the microscopes, so they got
insurance on them and required that I keep large plastic tarps to cover
them in case the sprinklers went off. They never did, howwever...
One Saturday, we had a huge rainstorm and the Dean's secretary called
me saying their seemed to be a "little problem in the lab." I rushed over
there to find that the drains at the top of the stairs had filled with
leaves and likewise the drain at the bottom and huge amounts of water,
like a waterfall, was cascading down the steps and filling up the lab.
The building power had gone off and the emergency lighting was on in the
hallways. I called for the maintenance crew, but they were priortized
to go save "expensive" student computers in another building. So there
I was with my broom, frantically cleaning drains and pushing water, feeling
like a character in Walt Disney's Fantasia. Because of this long and
intense fight, the only thing we lost was the carpeting, papers and books
on the lowest shelf. Saved the floor tile. The minimal rust on the feet of
the EM's wasn't even noticeable. The lab was fine. Sprinklers a non-issue.
The real devastation was done by the administration, later, when a
new Dean said "What are electron microscopes good for?" eyeing the space and
salary line he could use for a new Executive Secretary. But that's another
story. Kind regards and good wishes to all, Beverly



--Boundary (ID JlM655DZrjy0J4/C5dhX7g)--


----- End forwarded message






From: Gib Ahlstrand :      giba-at-puccini.crl.umn.edu
Date: Thu, 4 Apr 1996 13:38:04 -0600
Subject: Re: Water and the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In message {3163F416-at-hq_smtp} "Craig, Bob" writes:
}
} Good Morning All,
} Bob Citron's response to David Rothbard reminded me on an incident we had
} with an AMRAY 1000 many years ago (20?). A cooling water line on the
} diffusion pump failed and filled the Plexiglass housing of the high voltage
} supply with treated cooling water. Upon discovery me heart entered my
} throat and visions of disaster entered my mind. As in Bob's case, the
} service engineer calmly drained the water, rinsed the power supply with
} deionized water and dried it with a hair dryer. Other component also have
} been wet, which he dried with Kim Wipes. When he fired it up I expected
} sparks to fly, etc. No way!! The microscope worked like nothing had
} happened. Today the scope is 23 or 24 years old and still running! These
} experiences would indicate that older analog instruments were pretty hardy.
} How a modern digital instrument would react to such treatment is a good
} question.
} Bob Craig
} OSRAM SYLVANIA INC.
} Beverly, MA 01915
} craig-at-rd.sylvania.com


First of all, I'm very glad for Bob & Bob that water on their SEM's did not
cause any serious problems that simple drying out did not cure.

But I must relay my experience with a wet SEM lest folks get the impression that
these instuments are waterproof (no one would ever REALLY think that would
they?).

In past years, we have had water accidents in labs over our heads, in some cases
two floors above, and water has leaked down through cracks in their concrete
floors, through our ceiling, and onto our instruments (kind of like magnetic
atraction). In one case enough water got into the SEM so that a circuit board
that runs the Magnification readout and a few others were damaged a bit. Even
after replacing some transistors, the boards still blew out when powered up
again. My service engineer had to clean dirty residue left behind by the water
off the boards - many boards - with ethanol and a toothbrush to get rid of
ground paths causing electrical shorts. Naturally, such events are not covered
by service contracts, so we sent the bill upstairs to the offending lab. Some
one up there left water aspirator running all weekend, and naturally thats a
good time for paper towels to fall off the wall into the sink to plug up the
drain.

We then built plexiglass "deflectors" over our SEM and TEM, big sheets suspended
from the ceilings, to deflect any water coming down to the sides onto the floor.
Since installation in about 1985 they have saved us twice. We feel we got off
easy, as the potential was there for really serious big time damage.

The overhead panels also prevent some dust from falling onto the instruments.
And they make a great conversation piece when tours come through!

If there is any potential danger in your labs from overhead water pipes or water
from rooms above, I'd recommend that folks put up some kind of water sheding
panels. Could save a big head ache if the water ever comes down.


Gib Ahlstrand, MMS Newsletter Editor
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu

"MICROSCOPY & MICROANALYSIS 96" in Minneapolis, Minnesota, Aug.11-15, 1996





From: Shin-ichi Yamane :      a0c101u-at-cc.miyazaki-u.ac.jp
Date: Fri, 05 Apr 1996 15:08:06 +0900
Subject: Re: EM: Marine invertebrate texts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9604050608.AA00087-at-S_Yamane.cc.miyazaki-u.ac.jp}

Tue, 2 Apr 1996 13:20:10 +1100
richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood) wrote.
} Dear Microscopists,
} Does anyone out there know of a textbook or any other reference material
} which covers integumental ultrastructure - cuticle, epithelium and
} associated structures and also oogenesis - in marine invertebrates. We have
} a marine science student who is studying subclass Ascothoracida who is
} having problems getting the ultrastructure info she needs.
}
} Thanks in advance.



Richard

Do you know " Comparative Animal Cytology & Histology " written by
U.Welsch and V.Storch (1976), Sidgwick & Jackson Limited, London ?
This fundamental textbook translated from Germany contains "Chapter 3.
Integument" and "Chapter 11. Reproductive organs and cells".
Descriptions, of course, cover cuticle's ultrastructures of animals
of various phyla.
I think it's an available textbook for students though had been
issued two decades ago.


S. Yamane




----------------------------------------------------
Shin-ichi Yamane [ a0c101u-at-cc.miyazaki-u.ac.jp ]
Division of Marine Bioproduction
Faculty of Agriculture, Miyazaki University,
Gakuen Kibanadai Nishi 1-1, Miyazaki 889-21, Japan
Phone : 0985-58-2811 ext.3356
----------------------------------------------------




From: Beth Trend :      trend-at-cems.umn.edu
Date: Fri, 5 Apr 1996 11:42:39 -0600
Subject: Re: Water and the SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604051643.KAA22518-at-Sparc5.Microscopy.Com}

In message {199604041938.NAA17558-at-puccini.crl.umn.edu} "Gib Ahlstrand" writes:

} If there is any potential danger in your labs from overhead water pipes or
} water
} from rooms above, I'd recommend that folks put up some kind of water sheding
} panels. Could save a big head ache if the water ever comes down.
}


Sounds like a good idea ... but do you put them above or below the sprinkler
heads? :)


_______________________________________________________________________
Beth Trend trend-at-cems.umn.edu http://charfac.cie.umn.edu
Coordinator, Characterization Facility University of Minnesota
Center for Interfacial Engineering
100 Union St SE Room 187 phone 612-624-1365 fax 612-626-7530
Minneapolis, MN 55455







From: EmLab :      EMLAB-at-OPUS.MCO.EDU
Date: Fri, 05 Apr 1996 13:35:28 -0500 (EST)
Subject: Re: Sprinklers in the EM Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In my current place of employment there are no sprinklers in the lab.
I do agree with David that they are a mixed blessing. If they do open up, there
is the possibility of damage to epquiment but in the real case of a fire,
loss of epquiment is minor compared to a loss of a life.
In my preveiously employment, the maintance folks upgraded the sprinkler
system to a water sprinkler and removed the Halon system. Their reasoning was
that a person will get wet, but not dead as with the Halon system.
If a person is worried about electrocution with the water system, all
electricty should be able to be shut off from a remote electrical substation.

Best to all,
Ed Calomeni
Medical College of Ohio
Toledo, OH
emlab-at-opus.mco.edu




From: William Tivol :      tivol-at-wadsworth.org
Date: Fri, 5 Apr 1996 13:37:41 -0500 (EST)
Subject: Re: SEM - Surface Area or Surface Roughness Calculations

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


}
} Does anyone know of software that can calculate surface area or
} surface roughness parameters from a stereo pair obtained from an SEM?
}
Dear Dan,
If you want the true surface area--not just the projected area--
STERECON will allow you to reconstruct the volume, approximate the surface
with polygonal tiles and add these up to get an area. I don't know whether
it calculates roughness parameters, but these might be easy to determine
from the true and projected areas. I assume you are not concerned with
the fractional-dimensional properties which make the true area and the
roughness dependent on the length scale.
Yours,
Bill Tivol




From: Alan S. Pooley :      pooley-at-ahab.rutgers.edu
Date: Fri, 5 Apr 1996 15:08:47 -0500 (EST)
Subject: sprinklers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Our building (1992) was designed with a computer room and sem room next
to each other but independantly run and served by a single sprinkler
system where the pipe is filled with air until a fire is detected by
a heat sensor (or we open a valve in a red box like a fire alarm)
then a pump fills the pipe with water but the sprinkler still has to get
hot enough to melt to start water flowing (each room independant)

Email me direct if you need me to find out details from the planners


Alan S. Pooley ,PhD Bivalve shell SEM & shape analysis
SEM/Morphometrics lab
Marine & Coastal Sciences
Rutgers University
908 932 8959 ext 225
Pooley-at-ahab.rutgers.edu





From: Margaret H Malay :      malay-at-csd.uwm.edu
Date: Fri, 5 Apr 1996 11:17:52 -0600 (CST)
Subject: Midwest Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



*****************************************************
**************** !!! REMINDER !!! *****************
*****************************************************

Microscopy of Surfaces and Interfaces
Midwest Microscopy and Microanalysis Society Workshop
UWM Microscopy Open House

April 26, 1996

Please register by April 15, 1996
More information and online registration at
http://www.uwm.edu/Dept/LSS/meeting.html

PROGRAM
Golda Meir Library Conference Center
University of Wisconsin - Milwaukee

Opening Ceremonies
8:45 - 9:00

Morning Session

9:00 - 9:30
J. Mansfield
University of Michigan - Ann Arbor
"Materials Science and Biological Applications of Environmental Scanning
Electron Microscopy"

9:30-10:00
B. Tonner
University of Wisconsin - Milwaukee
"X-ray Microscopy for Chemical Analysis of Surfaces"

10:10 - 10:15 Coffee Break and Mounting of Posters
Sponsored by College of Letters and Sciences

10:15 - 10:45
T. Kelly
University of Wisconsin - Madison
"Three-Dimensional Atom Probe Microscopy"

10:45-11:15
J. Nogami
University of Wisconsin - Milwaukee
STM: "Studies of Metal Growth on Silicon Surfaces"


11:15-11:45
L. D. Marks
Northwestern University - Evanston
Surface HREM: "Old techniques on Old Surfaces - Surprising New Results"

Open House Lunch

11:45 - 1:00
Lunch sponsored by Nissei Sangyo America and Hitachi Scientific Instruments.
Poster Session

1:00 - 1:45
Poster viewing and poster competition
Poster prizes are sponsored by the MMMS.

Afternoon Session

1:45-2:15
J. M. Gibson
University of Illinois - Urbana
TEM: "All you ever wanted to know about building your own expensive and
complicated microscope, and then waiting five years until it finally works"

2:15 - 2:45
S. Babcock
University of Wisconsin - Madison
TEM/HREM: "Electron Microscopy Studies of Grain Boundaries in Bicrystals"

2:45 - 3:15
V. Dravid
Northwestern University - Evanston
"Analytical Electron Microscopy and Holography of Interfaces: Making a
Mountain out of a Mole Hill"

3:15 - 3:30 Coffee Break and Judging of Posters
Sponsored by Laboratory for Surface Studies

3:30 - 4:00
N. Browning
University of Illinois at Chicago
"High Angle Annular Dark Field in a Dedicated STEM"

4:00 - 4:30
K.L. Merkle
Argonne National Laboratory
"Relaxation Modes in Metal and Oxide Grain Boundaries"

4:30- 4:45
Poster Awards Ceremony and Closing Address
M. Gajdardziska-Josifovska, Workshop Organizer

Open House Tour

REGISTRATION FEES:
Registration is free for the members of the Midwest Microscopy and
Microanalysis Society. Members of the MicroBeam Analysis Society are members.
The workshop registration fees for non-members are same as the membership fees
and can be paid at the conference site (by check or cash):
Regular Member ($10)
Student Member ($5)

POSTERS
All lectures are by invitation only. Poster contributions by students,
researchers and microscope manufacturers are welcome. There are no size
restrictions for the posters, however they should be mounted on a sturdy
backing suitable for display on easels. The poster must be relevant for the
broad topic of microscopy of surfaces and interfaces.

POSTER COMPETITION

Graduate students are particularly encouraged to present their work and
they are eligible to enter a poster competition. One grand prize and two
first prizes will be given by the Midwest Microscopy and Microanalysis
Society. The grand prize must be used by the winner to attend a microscopy
related session at a major national/international conference.

Grand Prize: $500
1st Biological Sciences:$100
1st Physical Sciences: $100








From: Martin Kohler :      mk-at-enk.ks.se
Date: Sat, 06 Apr 1996 03:22:43 +0200
Subject: slide printers?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



High resolution slide printers?

I want to find out what slide printers (for 35mm film) there are on the
market with the following characteristics:
- generates sharp, high res, high quality slides
- has driver for Windows NT

Two apparently good slide printers are:
1. Polaroid Digital Palette HR 6000 (will have NT driver)
http://www.polaroid.com/homepage.htm
2. GALLERIA made by Mirus Industries (does not have NT drive)
http://www.mirus.com/

Could you please give any comments on these printers or suggest other slide
printers?

Thanks,
Martin

---------------
Martin K=F6hler
mk-at-enk.ks.se
---------------





From: John F. Mansfield :      jfmjfm-at-umich.edu
Date: Fri, 5 Apr 1996 19:24:48 -0500
Subject: SEM for sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1383385234.19715-at-mse.engin.umich.edu}

For Sale Hitachi S520 SEM.

30kV W-filament diffusion pumped microscope.
Secondary electron detector (Everhart Thornley).
BSE detector GW Electronics.
BSE detector Robinson.
S5679 Channelling Attachment.
S5010TV system.

Kevex 8000 XEDS system
10mm sq Be detector
needs new hard disk but otherwise operational.

Asking $23,000, but will consider offers.
Please contact me at the address below (email, phone, snail mail).


Jfm.


John Mansfield
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX (313)936-3352
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html






From: John F. Mansfield :      jfmjfm-at-umich.edu
Date: Sat, 6 Apr 1996 08:32:24 -0500
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {199604061622.LAA12559-at-IndyNet.indy.net}
To: DANIEL C VAN HART {vanhart-at-vnet.ibm.com} ,
Microscopy list {microscopy-at-Sparc5.Microscopy.Com}

Since this disease has had a fair degree of expsoure on the TV recently can
someone describe what it is and how it could be related to the BSE disease?

Please keep it in laymans terms if you can, I am a physicist not a physician!

Many thanks.
And sorry this isn't a Microscopy question.

Jfm.



John Mansfield
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (313)936-3352 FAX (313)936-3352
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html






From: Jay Jerome :      jjerome-at-bgsm.edu
Date: Sun, 7 Apr 1996 16:08:57 -0400 (EDT)
Subject: AREMS\SEMS 96 Meeting, Greenville SC

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




JOINT AREMS/SEMS MEETING
GREENVILLE, SC

WEDNESDAY

8:00 am - 4:00 pm REGISTRATION
12:30 pm-12:45 pm OPENING REMARKS
Dennis Barr, AREMS President-Elect
Jay Jerome, SEMS President

SESSION I

12:45 pm- 1:30 pm INVITED SPEAKER
Margaret Ann Goldstein
Baylor College of Medicine

"Imaging with microscopes and computers"

1:30 pm -2:30 pm ROUNDTABLE DISCUSSION
MODERATOR: Gene Michaels, Univ of GA

"Microscopy for kids of all ages"

2:30 pm -4:30 pm POSTER SESSION

Pennington SEM evaluation of the surface of an intra-arterial sensor
retrieved from critically ill pediatric patients

Johnsrude Intraerythrocytic inclusions associated with iridoviral
infection in a fer-de-lance (Bothrops moojeni) snake

Vaughn The effects of exogenous calcium oxalate crystals on
renal epithelial cells

Vaughn Electroreceptor organ development in larval electric eels
(Electrophorus electricus)

Gambling Remote printing of large format images

Whittaker Role of electron microscopy in diagnosis of the
transmission of fibropapillomas in Green turtles,
Chelonia mydas.
5:00 pm- 6:00 pm AREMS BUSINESS MEETING AND RECEPTION

2:30 pm -8:00 pm VENDOR SESSION

4:00 pm -8:00 pm VENDOR SPONSORED TUTORIALS


THURSDAY

SESSION II

9:00 am - 9:45 am INVITED TALK
Mike Kizer
Clinch Valley College
"Life Cycle Stages of H. diminuta by SEM & Freeze Cracking"


RUSKA STUDENT COMPETITION

Abstract
No.

9:45 1 Effects of recombinant bovine somatotropin (rbST) on
satellite cell and myofiber nuclei proliferation of beef
calves. R.C. Vann.

10:00 am -10:45 am INVITED TALK
Mark Teaford
Johns Hopkins
"Analysis of rates of tooth wear via scanning electron microscopy"

10:45-11:15 am COFFEE BREAK

SESSION III

11:15 am -12:00 pm INVITED TALK
Norman Herz
University of Georgia
"Innovative uses of SEM in the determination of weathering of statues"

12:00 pm - 1:30 pm LUNCH ON YOUR OWN

SESSION IV

1:30 pm - 2:15 pm INVITED TALK
Fred Stevie
AT&T
"Focused Ion Beam application in semiconductor applications"

2:15 2 Low kV analysis of uncoated and of coated artificial root
caries. DG Gantt

3:00 3 Computer-assisted analysis of radial symmetry in airway
epithelial cilia: a new perspective on assesment of congenital
ciliary defects. JL Carson.

3:15 4 Free radical derived oxidants and endothelial cell dysfunction
in a rat model of diabetic retinopathy. E Ann Ellis.

3:30 5 Fungi associated with heating, ventilating, and air
conditioning (HVAC) systems in the southeastern united states.
RB Simmons.

3:45 6 Negative stain electron microscopy (EM) of self-assembled
gastroenteric virus capsids. CD Humphrey.

4:00 7 Ultrastructure of conida and conidium germination in the plant
pathogenic fungus Alternaria cassiae. CW Mims.

4:15 pm - 4:30 pm COFFEE BREAK

4:30 pm - 5:15 pm INVITED TALK

Locke Christman
FEI Co.
"Focused ion beam secondary ion mass spectrometry (FIB SIMS)
expands the capabilities and the applications of focused ion beam
systems."


6:00 pm - 7:00 pm SOCIAL HOUR
7:00 pm- AREMS/SEMS AWARDS BANQUET
FEATURING AN AFTER DINNER TALK BY:
Mark Teaford
Johns Hopkins
"Making teeth talk"

FRIDAY


7:30 am - 9:00 am SEMS BUSINESS BREAKFAST

SESSION V

9:00 am - 9:45 am INVITED SPEAKER

Robert Price
University of South Carolina, Medical School
"Experimental models in the study of early heart development"

9:45 8 Microanalysis in environmental science: determination of the
micro-environment of pollutants and the mechanism of
remediation of contaminated soils. TB Vander Wood.

10:00-10:40 INVITED SPEAKER
Michael M. Thomason, Diane Petrovich, Steve Myles
Analytical Services Group, Simpsonville, SC
"SEM Analysis of nonwoven fabrics used in disposable diapers and other
converted products."

10:40 9 The quantification of structure within polycrystalline fiber.
KE Robinson.

10:55-11:10 COFFEE BREAK

11:10 - 11:50 INVITED SPEAKER
Rathna Perera
" The use of electron microscopy in high performance textiles"

11:50 10 Microscopic and chemical analyses of flax and flax retting. DE
Akin.

12:05-12:45 INVITED SPEAKER
T.J. Stark, D.P. Griffis, P.E. Russell
North Carolina State University, Raleigh, NC
"H2O Enhanced focused ion beam micromachining"

12:45 CLOSING REMARKS
Mark Farmer, SEMS President-Elect
John Herr, AREMS President


For further information contact:
Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-bgsm.edu *
**************************************************************





From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Sun, 7 Apr 1996 19:00:06 -0500
Subject: Re: Atlas of prokaryotes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Micro people: besides requests for copies of responses to my question, here
are the responses to the question "Is there a good 1990 or later prokaryote
SEM/TEM atlas?":

} A good virus morphology reference is 'Virus Morphology', CR Madeley and AM
} Field, 2nd ed, published by Churchill Livingston, Longman UK, (1988
} however) ISBN 0-443-02784-6.
} regards,
}
} Richard Easingwood
} South Campus Electron Microscope Unit
} School of Medical Sciences
} University of Otago
} PO Box 913
} Dunedin
} NEW ZEALAND

} From: "Jane A. Fagerland (847) 935-0104" {FAGERLAND.JANE-at-igate.pprd.abbott.com}
} Subject: Re: Atlas of prokaryotes
}
} Gee, maybe we should be collecting good examples and publish a book of
} our own! Thanks.

I seems that you're right Jane, at least for the bacteria and
archea. Maybe Madeley and Field are working on a 3rd edition?
But before anyone suggests that I do it, I'm a
creepy-crawlyologist. Anyone who works on prokaryotes would look at my name
and wonder who I am.
(Aside to John Bozzola: the authors of Brock et al. _The Biology of
Micro-organisms_ are all at Southern Illinois...)
Phil

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu






From: Scott Holt :      102467.2752-at-CompuServe.COM
Date: 08 Apr 96 09:40:29 EDT
Subject: TEM Epoxy: G1

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

Two weeks ago, I posted a message in response to questions about the composition
of GATAN's G1
TEM Epoxy. It has been brought to my attention, thankfully, that I have
mis-spoken, and should make
a corrective statement. My mistake was in assuring you that GATAN's G1 Epoxy
was the same as
EPOXY TECHNOLOGY's 353ND product. I should have stated that it was only my
personal observation
that the two are similar in characteristics, and that the MSDS sheets are
similar.

As GATAN has an interest in selling their G1 Epoxy, it was incorrect of me,
especially as we can, in the
extreme, be considered competitors, to connect their epoxy directly with EPO
TECH 353ND. However,
my purpose was only to direct you to a source for information on hazardous
components, as requested.
Please excuse my error.

MSDS information on EPO TECH 353ND can be obtained from EPOXY TECHNOLOGY by
calling:(508)667-3805.

Best Regards to all,
Scott D. Holt
BUEHLER
41 Waukegan Rd.
Lake Bluff, IL 60044
(847)295-4546





From: NORM OLSON :      NHO-at-bragg.bio.purdue.edu
Date: Mon, 8 Apr 1996 9:35:44 -0500 (EST)
Subject: Faraday cage holder

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a commercial source for a TEM specimen holder
that has a Faraday cage? We are looking for one for a Philips
EM420 and related Philips scopes. I know that Gatan, Inc has
them but we were also wondering if there were any other vendors.
We would also consider a used holder if one was available.

Thanks

Norm Olson
*******************************************
* Norm Olson *
* Department of Biological Sciences *
* Lilly Hall of Life Sciences *
* Purdue University *
* West Lafayette, IN 47907 *
* *
* Phone: 317-494-5643 *
* FAX: 317-496-1189 *
* email: nho-at-bragg.bio.purdue.edu *
* *
*******************************************





From: Jay Jerome :      jjerome-at-bgsm.edu
Date: Mon, 8 Apr 1996 10:44:50 -0400 (EDT)
Subject: 96 AREMS/SEMS meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



My apologies to those not interested in this meeting. I am resending this
announcement because in my first transmission I attached the wrong file.
This submission contains pertinent information about dates, registration
fees, etc. I knew I shouldn't do anything important on a Friday Afternoon.

JOINT AREMS/SEMS MEETING
GREENVILLE, SC
APRIL 24-26, 1996

LOCATION: Hyatt Regency, Greenville SC. 803-235-1234
Located in Downtown Greenville, SC
$82/night (single or double)

OTHER HOTELS WITHIN SHORT DRIVING DISTANCE:
Hampton Inn, 1-800-Hampton, 803-288-1200, $50-75
Holiday Inn, 803-297-6300, $50-75
La Quinta Inn, 1-800-531-5900, $25-50
Holiday Inn Express, 803-297-5353, $25-50

REGISTRATION FEE: $25 members/$35 non-members/$15 student (with ID) and
technicians
1 Day registration: $10

THURSDAY EVENING BANQUET: $22

You may register at the meeting. For arrangements information contact:
JoAn Hudson, EM Facillity, Clemson University, 803-656-2465

PROGRAM
WEDNESDAY

8:00 am - 4:00 pm REGISTRATION
12:30 pm-12:45 pm OPENING REMARKS
Dennis Barr, AREMS President-Elect
Jay Jerome, SEMS President

SESSION I

12:45 pm- 1:30 pm INVITED SPEAKER
Margaret Ann Goldstein
Baylor College of Medicine

"Imaging with microscopes and computers"

1:30 pm -2:30 pm ROUNDTABLE DISCUSSION
MODERATOR: Gene Michaels, Univ of GA

"Microscopy for kids of all ages"

2:30 pm -4:30 pm POSTER SESSION

Pennington SEM evaluation of the surface of an intra-arterial
sensor
retrieved from critically ill pediatric patients

Johnsrude Intraerythrocytic inclusions associated with iridoviral
infection in a fer-de-lance (Bothrops moojeni) snake

Vaughn The effects of exogenous calcium oxalate crystals on
renal epithelial cells

Vaughn Electroreceptor organ development in larval electric
eels
(Electrophorus electricus)

Gambling Remote printing of large format images

Whittaker Role of electron microscopy in diagnosis of the
transmission of fibropapillomas in Green turtles,
Chelonia mydas.
5:00 pm- 6:00 pm AREMS BUSINESS MEETING AND RECEPTION

2:30 pm -8:00 pm VENDOR SESSION

4:00 pm -8:00 pm VENDOR SPONSORED TUTORIALS


THURSDAY

SESSION II

9:00 am - 9:45 am INVITED TALK
Mike Kizer
Clinch Valley College
"Life Cycle Stages of H. diminuta by SEM & Freeze Cracking"


RUSKA STUDENT COMPETITION

Abstract
No.

9:45 1 Effects of recombinant bovine somatotropin (rbST) on
satellite cell and myofiber nuclei proliferation of beef
calves. R.C. Vann.

10:00 am -10:45 am INVITED TALK
Mark Teaford
Johns Hopkins
"Analysis of rates of tooth wear via scanning electron microscopy"

10:45-11:15 am COFFEE BREAK

SESSION III

11:15 am -12:00 pm INVITED TALK
Norman Herz
University of Georgia
"Innovative uses of SEM in the determination of weathering of statues"

12:00 pm - 1:30 pm LUNCH ON YOUR OWN

SESSION IV

1:30 pm - 2:15 pm INVITED TALK
Fred Stevie
AT&T
"Focused Ion Beam application in semiconductor applications"

2:15 2 Low kV analysis of uncoated and of coated artificial root
caries. DG Gantt

3:00 3 Computer-assisted analysis of radial symmetry in airway
epithelial cilia: a new perspective on assesment of
congenital
ciliary defects. JL Carson.

3:15 4 Free radical derived oxidants and endothelial cell
dysfunction
in a rat model of diabetic retinopathy. E Ann Ellis.

3:30 5 Fungi associated with heating, ventilating, and air
conditioning (HVAC) systems in the southeastern united
states.
RB Simmons.

3:45 6 Negative stain electron microscopy (EM) of self-assembled
gastroenteric virus capsids. CD Humphrey.

4:00 7 Ultrastructure of conida and conidium germination in the
plant
pathogenic fungus Alternaria cassiae. CW Mims.

4:15 pm - 4:30 pm COFFEE BREAK

4:30 pm - 5:15 pm INVITED TALK

Locke Christman
FEI Co.
"Focused ion beam secondary ion mass spectrometry (FIB SIMS)
expands the capabilities and the applications of focused ion beam
systems."


6:00 pm - 7:00 pm SOCIAL HOUR
7:00 pm- AREMS/SEMS AWARDS BANQUET
FEATURING AN AFTER DINNER TALK BY:
Mark Teaford
Johns Hopkins
"Making teeth talk"

FRIDAY


7:30 am - 9:00 am SEMS BUSINESS BREAKFAST

SESSION V

9:00 am - 9:45 am INVITED SPEAKER

Robert Price
University of South Carolina, Medical School
"Experimental models in the study of early heart development"

9:45 8 Microanalysis in environmental science: determination of the
micro-environment of pollutants and the mechanism of
remediation of contaminated soils. TB Vander Wood.

10:00-10:40 INVITED SPEAKER
Michael M. Thomason, Diane Petrovich, Steve Myles
Analytical Services Group, Simpsonville, SC
"SEM Analysis of nonwoven fabrics used in disposable diapers and other
converted products."

10:40 9 The quantification of structure within polycrystalline fiber.
KE Robinson.

10:55-11:10 COFFEE BREAK

11:10 - 11:50 INVITED SPEAKER
Rathna Perera
" The use of electron microscopy in high performance textiles"

11:50 10 Microscopic and chemical analyses of flax and flax retting.
DE
Akin.

12:05-12:45 INVITED SPEAKER
T.J. Stark, D.P. Griffis, P.E. Russell
North Carolina State University, Raleigh, NC
"H2O Enhanced focused ion beam micromachining"

12:45 CLOSING REMARKS
Mark Farmer, SEMS President-Elect
John Herr, AREMS President




Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-bgsm.edu *
**************************************************************





From: akracher-at-iastate.edu (Alfred Kracher)
Date: Mon, 8 Apr 1996 10:04:06 -0600
Subject: Re: Creutzfeld-Jakob disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: akracher-at-pop-1.iastate.edu
Message-Id: {v02130501ad8ee8f70559-at-[129.186.121.151]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

The article "The Prion Diseases" by Stanley B. Prusiner appeared in
Scientific American of January 1995 (vol. 272 no. 1), p.48.

Alfred Kracher
akracher-at-iastate.edu
http://www.public.iastate.edu/~akracher






From: Jane A. Fagerland (847) 935-0104 :      FAGERLAND.JANE-at-igate.pprd.abbott.com
Date: Mon, 08 Apr 1996 11:05:00 -0600 (CST)
Subject: Re: Creutzfeld-Jakob Disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta RANDD; Relayed; Mon, 08 Apr 1996 12:28:42 -0600
Mr-Received: by mta MCM$RAND; Relayed; Mon, 08 Apr 1996 12:28:45 -0600
Mr-Received: by mta RANDD; Relayed; Mon, 08 Apr 1996 12:28:55 -0600
Alternate-Recipient: prohibited
Disclose-Recipients: prohibited
Content-Return: prohibited

Here's a simplified version of the mad cow disease story:

The connection between BSE and C-J Disease is that both are thought to
be caused by prions, which are self-replicating, abnormally folded,
mutated versions of a normal cellular protein. There are similar
diseases in other species, the most notable being scrapie in sheep.
The diseases have relatively long incubation periods (years) and result
in a characteristic "spongiform" lesion in the brain. The diseases are
easily diagnosed post-mortem by the presence of little holes in the
brain, which can be detected with light microscopy. Prion particles
themselves can be visualized with negative staining and TEM.

Great Britain has had a huge problem with BSE beginning in the late
1980's. The problem in beef was caused by feeding scrapie-infected
sheep scraps to cattle, and there is good evidence that the disease
jumped from sheep to cattle. The fear is that if it could make that
jump, then maybe it could jump from cattle to humans. No one has
thought to wonder, if that were the case, why it hasn't jumped from
sheep to humans, since mutton is commonly eaten in Great Britain; but
this story is not about logic - it's about sensationalism.

C-J Disease sporadically and spontaneously occurs in about 1 in a
million humans. (There is another human prion disease called kuru that
is transmitted by eating human brains - the solution to that health
problem is pretty straightforward!) Until a few weeks ago, the British
health department has been adamantly sticking to the story that there
was no evidence for an increased incidence of C-J corresponding with the
current epidemic of BSE in British cattle. However, they recently
wavered and said they couldn't rule out a BSE connection to 10 cases of
a variant form of C-J that appears to have a shorter than usual
incubation period. The British press has had a heyday with that, and I
noticed that even the Chicago Tribune had a headline in yesterday's
paper stating that an Italian had died of mad cow disease. It's not
likely that the disease would be transmitted by eating beef from
infected cattle, particularly because prions reside in central nervous
system tissues (brain and spinal cord), and humans don't normally eat
those.

The problem in Britain right now is the proverbial tempest in a teapot.
The sad part is that many farmers are losing their livelihoods because
the British press was irresponsible in the way it reported the
information from the health department.

Enough said, I guess, before I climb any higher on the soapbox!

Jane A. Fagerland, Ph.D.
Dept. of Microscopy and Microanalysis
Abbott Laboratories
Abbott Park, IL 60064






From: William Tivol :      tivol-at-wadsworth.org
Date: Mon, 8 Apr 1996 17:30:24 -0500 (EDT)
Subject: Re: NonMagAlloy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Wil,

} I am trying to design a special specimen holder for a high resolution TEM,
} and am looking for a good alloy to use for this purpose; [snip]
} one that will have sufficient strength to withstand the forces [snip],
} yet is reasonably easy to machine.
}
We have had very good luck with aluminum, believe it or not. The
machinists like it too. In the HVEM, the specimen stage is held against
the translation apparatus by 1 at. applied over ~9/16" dia., and this seems
not to affect the resolution, etc. Unless you need exact dimensions be-
tween a fiducial point on the stage and (say) the center of the specimen,
you should be able to use almost any material. If you do need the stage to
have constant dimensions to an angstrom or so, how will you mount the
specimen to this accuracy?

} I think I have heard that some special grades of stainless steel, and
} possibly some heat-resistant alloys, are suitable for use in such
} applications. Does anyone know what alloys might be candidates, or what
} limiting values of magnetic properties (permeability, susceptibility, etc.)
} an alloy should have?
}
Be very careful with stainless steels. The machining process can
convert a non-magnetic form to one with magnetism. We had the shop make
a pin for a device which ended up ~1.5" from the beam, and there was a
magnetism problem which went away when the pin was replaced by one made
of phosphor bronze--another alloy to consider if aluminum is not suitable.
Yours,
Bill Tivol




From: Mary-ann Miller - ELCE/W95 :      m8miller-at-acs.ryerson.ca
Date: Mon, 8 Apr 1996 19:24:09 -0400 (EDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Please unsubscribe. Thank you





From: Mary-ann Miller - ELCE/W95 :      m8miller-at-acs.ryerson.ca
Date: Mon, 8 Apr 1996 19:26:26 -0400 (EDT)
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please Unsubscribe





From: Mary-ann Miller - ELCE/W95 :      m8miller-at-acs.ryerson.ca
Date: Mon, 8 Apr 1996 19:25:50 -0400 (EDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



unsubscribe





From: William Tivol :      tivol-at-wadsworth.org
Date: Tue, 9 Apr 1996 10:37:58 -0400 (EDT)
Subject: STERECON

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} can you please let me know how to obtain more information about
} STERECON. Is this shareware or a commercial package?
}
Dear Hasso, et al.,
For info on STERECON, check out J. Stru. Biol. v116, pp 93-98
(1996), M. Marko & A Leith, which describes the program and its capa-
bilities, or check our Web site www.wadsworth.org under SPIDER then
STERECON. For further info, please ask the real expert, Mike Marko,
at "marko-at-wadsworth.org". He would know the particulars of the cost
of obtaining STERECON as well as any technical info.
Yours,
Bill Tivol




From: Owen P. Mills :      opmills-at-mtu.edu
Date: Tue, 9 Apr 1996 12:53:27 -0400
Subject: pulse processor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi;

Can anyone confirm that a Kevex pulse processor (from a 7000 system) can be
used with a EDAX detector (with 184 pre-amp)?

Thanks,

Owen
Owen P. Mills
Michigan Technological University
Metallurgical & Materials Engineering
Rm 512 MME Building
Houghton, MI 49931
906-487-2002
906-487-2934 FAX
opmills-at-mtu.edu





From: dshubito-at-d.imap.itd.umich.edu (Dennis Shubitowski)
Date: Tue, 9 Apr 1996 14:50:45 -0400
Subject: Hi Res TEM Cameras

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02140b00ad905ffbe91a-at-[141.211.157.94]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I am looking for information on high resolution cameras that would
attach to a Philips CM100. I am familiar with Kodak's Megaplus 1.6
camera but was wondering what other manufacturers had to offer or
if Kodak has come out with anything more recent than the Megaplus.

Thanks,

Dennis Shubitowski
University of Michigan
School of Dentistry
dshubito-at-umich.edu






From: KINGSLAND, Arlene :      KINGSLAND-at-paprican.ca
Date: Tue, 9 Apr 1996 16:13:13 EST5EDT
Subject: Nanoplast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {MAILQUEUE-101.960409161312.416-at-pap386.paprican.ca}

Hello/Bonjour

I am about to learn a technique which is very new to me. I will
be embedding pulp fibres in nanoplast. I have been warned that this
will tend to be very brittle for trimming.
Does anyone have any experience with this, specifically with
cellulose fibres? Is there a way to prevent the brittleness?
Any general advice...I am green at this.

Thanks, Arlene Kingsland
Pulp and Paper Research Institute of Canada






From: vickie-at-MACC.WISC.EDU
Date: Tue, 9 Apr 1996 14:19:01 -0600
Subject: Symposium/Workshop Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Symposium on:

"Integrated Microscopy"

September 20 to 22, 1996

Co-hosted by:

Integrated Microscopy Resource (IMR)
at the University of Wisconsin-Madison
an NIH Biomedical Research Resource

and

Center for Fluorescence Imaging and Biotechnology
at Carnegie-Mellon University
an NFS Biotechnology Resource

Location:

The Wisconsin Center
702 Langdon Street
Madison, WI 53706


#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*

Presentations will focus on biological problems for which a combination of
microscopies [i.e. integrated microscopy] has been used. The speakers will
demonstrate by example the power, potential and limitations of various
microscopical techniques.

#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*

This year we are happy to invite attendees to present their work during the
Saturday evening poster session. There is a limit of approx. 40 posters
which can be displayed, therefore they will be accepted on a first come
first serve basis. A one page (8.5X11) typed abstract should accompany the
registration form and fees.

#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*#*

SCHEDULE

Friday Evening, September 20, 1996
5:00- 6:55 Reception and Poster Set-up
7:00- 7:45 Alan Waggoner - Amersham Life Science/Carnegie
Mellon University
7:50- 8:35 Andrew Dixon - Biorad Microsciences
8:40- 9:25 Fred Lanni - Carnegie-Mellon University

Saturday, September 21, 1996

8:00- 8:25 Hospitality Buffet
8:30- 9:15 David Mastronarde - University of Colorado
9:20-10:05 John Z. Kiss - Miami University of Ohio
10:10-10:40 Coffee Break
10:45-11:30 Mark Ellisman - Univ. California-San Diego
11:35-12:20 Patricia Calarco - Univ. California-San Francisco
12:25- 1:25 Lunch (on your own)
1:30- 2:15 Jeff Hardin - University of Wiscosin-Madison
2:20- 3:05 Jim G. McNally - Washington University
3:10- 3:55 Enrico Gratton - Univ. of Illinois Urbana/Champaigne

4:00- 6:00 Poster Presentations
6:00- 7:00 Exhibitor's Show
7:00- 9:00 Buffet Dinner

Sunday, September 22, 1996

8:30- 8:55 Hospitality Buffet
9:00- 9:45 Steven M. Block - Princeton University
9:50-10:35 Conly Rieder - Wadsworth Center for Labs/Research
10:40-11:10 Coffee Break
11:15-12:00 Scott E. Fraser - California Institute of Technology
12:05-12:50 Phillip G. Haydon - Iowa State University

1:00- 2:00 IMR Tour

FEES:
General Registration $ 80.00
(Includes: Opening Reception, Social and Buffet Dinner, Coffee
Breaks and Materials)

Student and Local Registration $ 50.00
(Includes: Opening Reception, Social, Coffee Breaks and

Materials)

Abstract Handling Fee $ 25.00

Dinner Ticket $ 20.00


fFOR ADDITIONAL INFORMATION and PROGRAM UPDATES CONSULT OUR WEB SITE:

http://www.bocklabs.wisc.edu/imr/imr.html


TO RECEIVE A BROCHURE AND REGISTRATION FORM

WRITE: IMR
Univ. Wisconsin-Madison
1675 Observatory Drive
Madison, WI 53706

OR EMAIL: imradmin-at-calshp.cals.wisc.edu


********************************************************************************

Following the symposium, the IMR will be conducting a 2-day workshop. We
will be presenting lectures and provide "hands-on" experience for the
following techniques:
* Multiple-photon excitation imaging
* 4D DIC imaging
* Cryo-SEM
* High pressure freezing
* Reversible embeddment for SEM and TEM

Workshop attendence will be limited to 25 participants. A letter of
application is required. Once accepted a fee of $150.00 will be due.

********************************************************************************





From: Frantisek Weyda :      weyda-at-entu.cas.cz
Date: Wed, 10 Apr 1996 10:36:40 -0900 (PDT)
Subject: Re: Nanoplast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Arlene Kingsland
Pulp and Paper Research Institute of Canada


Dear Arlene,

concerning nanoplast resins I have no experience with cellulose fibres
but I would like to recommend you my paper:

Weyda F., 1990:
Notes on the use of Nanoplast FB 101 for transmission
electron microscopy.
J.Electron Microsc.Tech., 16: 356-357

where you could find some useful information about the procedure of
embedding into Nanoplast FB 101.

Sincerely,

RNDr.Frantisek Weyda,CSc. | Frantisek Weyda, PhD.
Entomologicky ustav AV CR | Institute of Entomology
Branisovska 31
370 05 Ceske Budejovice
Ceska republika | Czech Republic
---------------------------------------------------------------------
tel.: (38) 817 linka 257 | phone: +42 38 817 ext.257
fax.: (38) 43625 | Fax.: +42 38 43625

Internet e-mail: weyda-at-entu.cas.cz







From: Karl Johnson :      kjohnson-at-haverford.edu
Date: Wed, 10 Apr 1996 10:16:49 +0000
Subject: Teaching Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {316B8A91.18CC-at-haverford.edu}

The Department of Molecular, Cellular and Developmental Biology at
Haverford College seeks an individual to co-teach a 7 week intensive
course in microscopic technique during the first quarter of
1996-1997 academic year. Part of the major core, this laboratory
has an expected enrollment of between 30 and 40 students and meets
in two sections over five afternoons a week. This laboratory
experience should include hands-on student experience in light,
fluorescence and electron microscopy with a focus upon cell and
tissue structure and function. Equipment used in the course
includes two Nikon Labophot fluorescent microscopes and an Hitachi
H-600 EM. The ideal candidate would have a Ph.D. (although A.B.D.
will be considered) and experience in electron microscopy who is
interested in gaining teaching experience in a liberal arts
environment.

The department is also seeking an individual to co-teach in our
introductory course "The Cellular Basis of Life" in the Spring
semester.

Interested candidates for either (or both) of these positions should
mail or fax a cover letter and curriculum vitae as soon as possible
to:

Slavica Matacic, Ph.D. or Karl Johnson, Ph.D.
smatacic-at-haverford.edu kjohnson-at-haverford.edu
(T) 610-896-1306 (T) 610-896-1305
Department of Molecular, Cellular and Developmental Biology
Haverford College 370 Lancaster Avenue
Haverford, PA 19041
(F) 610 896-4963

Additional information about the department and its program is
available via the World Wide Web at
http:www.haverford.edu/biology/welcome.html.

Haverford College is an affirmative action, equal opportunity
employer.




From: Jane A. Fagerland (847) 935-0104 :      FAGERLAND.JANE-at-igate.pprd.abbott.com
Date: Wed, 10 Apr 1996 09:18:00 -0600 (CST)
Subject: BSE and C-J Disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta RANDB; Relayed; Wed, 10 Apr 1996 09:38:55 -0600
Mr-Received: by mta MCM$RAND; Relayed; Wed, 10 Apr 1996 09:38:57 -0600
Mr-Received: by mta RANDC; Relayed; Wed, 10 Apr 1996 09:39:12 -0600
Alternate-Recipient: prohibited
Disclose-Recipients: prohibited
Content-Return: prohibited

Thanks for all the comments concerning my remarks about prion diseases.
I stand corrected on my assertion that humans don't usually eat central
nervous system tissue from sheep and cattle. Several folks from the UK
enlightened me - sheep and calf brains are considered a delicacy in some
areas of the UK and are also used as fillers in hamburger and sausage.
Also, soups are made from animals parts that may still have CNS tissues
attached to them.

Someone also pointed out that it was initially thought that scrapie
could not jump from sheep to cattle; thus, there's precedent for inter-
animal transmission, regardless of what seems likely right now.

For anyone wanting more information on BSE, there is now a BSE group.
To subscribe, send an e-mail to {majordomo-at-info.aphis.usda.gov} . Leave
the subject line blank, and type {subscribe bse} in the message. There
is also a BSE news and information service on the Web at
http://www.cabi.org/


Jane A. Fagerland, Ph.D.
Dept. Microscopy and Microanalysis
Abbott Laboratories
Abbott Park IL 60064






From: sbarlow-at-sunstroke.sdsu.edu (Steve Barlow)
Date: Wed, 10 Apr 1996 10:17:11 -0800
Subject: SEM:acc voltage test sample

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02110101ad91a36cfabc-at-[130.191.238.149]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

To all:

Two related questions:

I) is there any interest in compiling a list of samples that can be used in
lab classes to readily show different principles of microscope operation?
One example of one of my test samples would be .4 um nucleopore filters
with their nice round pores for demonstrating astigmatism and how to
correct for it. Are others willing to share their ideas in this
regard--I'll be happy to compile and post a summary

II) Along these same lines, does anyone have a suggestion for a test
sample to use in my SEM laboratory class that clearly demonstrate the
effects of changing accelerating voltage? I haven't come across an easily
obtainable, reproducible sample to clearly show:

1) enhancement of surface detail at low voltage vs high votage

2) ability to view uncoated samples at low voltage without charging vs high
voltage with charging

any suggestions in this regard would be welcome

steve

Dr. Steven Barlow
EM Facility/Biology Dept.
San Diego State University
San Diego, CA 92182-4614
phone: (619)594-4523
fax:(619)594-5676
email:sbarlow-at-sunstroke.sdsu.edu






From: slocombe-at-wsunix.wsu.edu (Peter Slocombe)
Date: Wed, 10 Apr 1996 12:48:01 -0700
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9604101842.AA03291-at-super.mhv.net}
To: nestor {Microscopy-at-Sparc5.Microscopy.Com}

unsubscribe





From: Stefan.Gunnarsson-at-devbiol.uu.se (Stefan Gunnarsson)
Date: Wed, 10 Apr 1996 21:23:13 +0100
Subject: Re: Stereology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: zoogun-at-strix.udac.uu.se (Unverified)
Message-Id: {v01540b00ad91c7db9c2d-at-[130.238.188.129]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Sorry for being a bit after the thread with this, but it would be nice to
know if there is any freeware stereology programs around. I don't need
anything very fancy but just something simple to use mainly for
demonstrations and teaching purposes.

TIA

Stefan

Stefan Gunnarsson
Microscopy & Imaging Unit, Uppsala University
Norbyvagen 18A
S-75236 Uppsala, Sweden
tel. +46 18 182638, fax. +46 18 182683






From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Wed, 10 Apr 1996 10:39:30 -0600
Subject: Stereo Imaging: Anaglyphs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130501ad9193550782-at-[131.230.97.69]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I am looking for red/green glasses for viewing some anaglyphs that I have
on slides. Ted Pella sells red/blue but I don't think they will work with
my pre-existing slides.

Also, why the change from red/green to red/blue? Is this due to the "color
bombardment" phenomenon that causes a strain to some viewers or are there
other reasons?

Any newer published references re anaglyphs besides the 1980 article by
Barber and Emerson in Scanning 3:202-206?

Many thanks.


#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Wed, 10 Apr 1996 10:39:30 -0600
Subject: Stereo Imaging: Anaglyphs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130501ad9193550782-at-[131.230.97.69]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I am looking for red/green glasses for viewing some anaglyphs that I have
on slides. Ted Pella sells red/blue but I don't think they will work with
my pre-existing slides.

Also, why the change from red/green to red/blue? Is this due to the "color
bombardment" phenomenon that causes a strain to some viewers or are there
other reasons?

Any newer published references re anaglyphs besides the 1980 article by
Barber and Emerson in Scanning 3:202-206?

Many thanks.


#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: James R. Stets :      stetsjr-at-ttown.apci.com
Date: Wed, 10 Apr 1996 15:17:57 -0400 (EDT)
Subject: Re:SEM Acc. Voltage Test Sample

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On April 10, Dr. Steven Barlow asked:

(text deleted)...does anyone have a suggestion for a test
sample to use in my SEM laboratory class that clearly demonstrate the
effects of changing accelerating voltage?

A few years ago, Dr. David Joy wrote an article which appeared in some
Hitachi literature about the advantages of low-voltage imaging. He used
a common carbon-film covered TEM grid imaged at 1 or 2 kV and the same
view at 20 kV. The low-voltage image clearly showed the carbon film,
while at 20 kV, the dominant feature was the grid bars, and the carbon
film seems to disappear.

The images were obtained at low magnification (less than X1000) so you
can do this even if you don't have a FESEM.

Jim Stets
Air Products and Chemicals, Inc.
Allentown, PA
stetsjr-at-ttown.apci.com
My opinions are my own, not my employer's






From: MelanieOwl-at-aol.com
Date: Wed, 10 Apr 1996 21:06:57 -0400
Subject: Re: Safety & Liquid N2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Just another late word on the LN2 issue. There are crogenic gloves available
which will properly protect your hands when handling liquid nitrogen or
shutting off frozen valves and handling frozen hoses. Although I agree that
sometimes safety issues don't necessary reflect common sense, having the
gloves doesn't hurt.

By the way, it snowed yesterday, again....when will we see Spring in the
Northeast?

Regards,
Melanie Behrens
Texaco Fuels and Lubricants Research
Beacon, NY
behrema -at- Texaco.com
914-838-7261




From: Robin Griffin :      rgriffin-at-eng.uab.edu
Date: Wed, 10 Apr 1996 10:12:20 -0500
Subject: SEM analysis of collagen in cartilage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {c=US%a=_%p=UAB%l=ENGEM0-960410151220Z-4934-at-engem0.eng.uab.edu}

HELP!

I have a biomedical engineering student that would like to analyze the
spatial arrangement of collagen arrangement in cartilage. I'm a
materials person and have absolutely no idea how to start, how to
prepare the samples, where to look up information or if it is even
possible. The papers we have show TEM analysis but we would prefer to
start with SEM.

Can anyone out there point me in the right direction.

Robin Griffin
Electron Optics Laboratory Manager
Materials and Mechanical Engineering
The University of Alabama at Birmingham




From: Walt Bobrowski (313) 996-7814 :      bobroww-at-aa.wl.com
Date: Thu, 11 Apr 1996 08:22:45 -0400 (EDT)
Subject: Re: Stereo Imaging: Anaglyphs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

MSA MICROSCOPY MAILING LIST {MICROSCOPY-at-Sparc5.Microscopy.Com}
Message-Id: {3545220811041996/A24505/SRVR05/11A45A162C00*-at-MHS}
X-Envelope-To: MICROSCOPY-at-MSA.MICROSCOPY.COM, bozzola-at-siu.edu
Autoforwarded: false
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; CHARSET=US-ASCII
Content-Transfer-Encoding: 7BIT
Importance: normal
Priority: normal
Sensitivity: Company-Confidential
Ua-Content-Id: 11A45A162C00
X400-Mts-Identifier: [;3545220811041996/A24505/SRVR05]
Hop-Count: 2

Red/Green glasses for viewing stereo images can obtained FREE from the Philips
Electron Optics web site -at- HTTP://www.peo.philips.com

Look in the subsection of 3-D images for the electronic order form.






From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Thu, 11 Apr 1996 08:26:22 GMT
Subject: Re: Stereo Imaging: Anaglyphs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 10:39 AM 4/10/96 -0600, you wrote:
} I am looking for red/green glasses for viewing some anaglyphs that I have
} on slides. Ted Pella sells red/blue but I don't think they will work with
} my pre-existing slides.
}
} Also, why the change from red/green to red/blue? Is this due to the "color
} bombardment" phenomenon that causes a strain to some viewers or are there
} other reasons?
}
} Any newer published references re anaglyphs besides the 1980 article by
} Barber and Emerson in Scanning 3:202-206?
}


Hi John. This question came up recently and I archived the responses
on our web page. Go to the address at the end of this message and click on
the "Tips & Tricks Wizard". You should find two links that might be useful.

1." Stereo Glasses, Red/ Green"
2. "Presentation of Stereo Pairs"

If you do not have web acess or cannot read the files, please let me
know and I will be happy to e-mail, fax, etc... them to you.

Hope this helps.




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {

Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 904-392-1295
ICBR EM Core Lab fax 904-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/





From: David Garrett :      DGARRETT-at-gab.unt.edu
Date: Thu, 11 Apr 1996 10:39:09 CST6CDT
Subject: MSA printer test file

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can someone remind me where the standard printer test file is
located. This is the file that was being printed on different
printers and then brought to the annual meetings for comparison.
Thanks,

***************************************
David Garrett "DGARRETT-at-GAB.UNT.EDU"
University of North Texas
Dept. Biological Sciences
(817)565-3964 Fax (817)565-4136
***************************************




From: keller-at-boulder.nist.gov (Bob Keller)
Date: Thu, 11 Apr 1996 08:38:03 -0600
Subject: SEM:acc voltage test sample

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: keller-at-arc1.mrd.bldrdoc.gov
Message-Id: {v01540b00ad92c6e3ab9b-at-[132.163.192.156]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Steve,

in regard to: 1) enhancement of surface detail at low voltage vs high voltage:

Try imaging a very ductile fracture surface (e.g. mild steel, etc.) for
demonstrating high vs. low keV effects. At the higher energy, the image
looks fine, but you get excessive intensity from thinner edges. At the
lower energy, that excess brightness is gone and you start to see fine
features in the edges as well as on (what you thought were) flatter
regions. The voltage effect is quite strong for comparisons between e.g.
25 keV vs. 5 keV.

in regard to: 2) ability to view uncoated samples at low voltage without
charging vs high
voltage with charging:

For viewing uncoated samples, try a ceramic, even a piece of glass. You
can get decent, uncharged imaging on SiO2 at around 1.5 to 2.0 keV. Above
that, all the charging you could want and then some.

Bob Keller
NIST
Materials Reliability Division
Boulder, CO






From: ldm3-at-apollo.numis.nwu.edu (L. D. Marks)
Date: Thu, 11 Apr 1996 14:39:04 -0500
Subject: Postdoctoral Position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

ajayan-at-hrem.mpi-stuttgart.mpg.de, aking-at-boundaries.eng.sunysb.edu,
alan_spargo-at-muwayf.uimelb.edu, b.d.hall-at-massey.ac.nz,
BABCOCK-at-coeadm.engr.wisc.edu, batson-at-watson.ibm.com, colliex-at-frsolii,
dave-at-physics.att.com, del-at-sol1.lrsm.upenn.edu,
differ-at-apollo.numis.nwu.edu, dvd-at-ruca.ua.ac.be, fmross-at-Csa2.LBL.Gov,
frchen-at-MSC.nthu.edu.tw, furukaz-at-nrim.go.jp, hobbs-at-MOLOKAI.MIT.EDU,
iijima-at-tgn.cl.nec.co.jp, john_spence-at-macmail.inre.asu.edu,
johnf-at-chem.gla.ac.uk, JSILCOX-at-msc.cornell.edu,
jws%uk.ac.bristol.pva.BITNET-at-anlvm.ctd.anl.gov, larsk-at-inorg.su.se,
maok-at-lbl.gov, mgj-at-csd.uwm.edu, MICROSCOPY-at-sparc5.microscopy.com,
mmdisko-at-erenj.com, nanotem-at-indirect.com, pabuffat-at-i2msg1.epfl.ch,
RAVEAU-at-frcpn11.in2p3.fr, Rebecca_Ai-RP3478-at-email.sps.mot.com,
rgronsky-at-garnet.berkeley.edu, Roar_Kilaas-at-macmail.lbl.gov,
SXU-at-PTD.intel.com, turner_p-at-chem.usyd.edu.au,
userrfe2-at-mts.ucs.ualberta.ca, van_tendeloo-at-ruca.ua.ac.be,
wos1-at-cam.ac.uk

Postdoctoral Position

A postdoctoral position will be available at Northwestern
University starting around August-October 1996. The position
will involve extensive use of HREM coupled with diffraction/image
simulations and crystallography on high temperature superconductors.
A strong background in electron microscopy and crystallography
is required, and strengths in the use of computers will be
important. The ability to work with others will also be very
significant, since the position will involve extensive collaborative
work.

Send CV plus the names of 3 referees with (seperately) letters
to:
L. D. Marks
Department of Materials Science
Northwestern University
Evanston, IL 60208, USA
http://risc1.numis.nwu.edu
ldm3-at-apollo.numis.nwu.edu




From: Dr. L. P. Stoter :      LPS-at-teknesis.demon.co.uk
Date: Thu, 11 Apr 1996 21:44:43 +0000
Subject: Microscopy and Analysis Services in USA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: (Unverified)
Message-Id: {v01510100ad932bde387f-at-[158.152.199.245]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Microscopy & Analysis Services in the USA


"Microscopy and Analysis" is published 6 times a year, in UK, European and
USA editions. It is widely circulated among all types of user, both
academic and commercial, in the areas of microscopy and associated
analytical techniques.

For a number of years, as a service to its readers, Microscopy and Analysis
has published in its UK edition a listing of laboratories selling their
services. The listing provides full contact details, plus a tabulation by
technique of the services offered by each laboratory, with a note on any
special capabilities or expertise.

This year, we plan to extend this feature to the USA edition. Laboratories
are invited to send in details of the services offered. Please send on a
single page, the following information:

1. Contact name, full postal address, telephone and fax numbers, and e-mail
address.
2. Indicate if your experience is in the materials or biological sciences,
or both.
3. List the instrumental techniques available in your laboratory, for
example TEM, light microscopy, EDX, Raman spectroscopy, etc.
4. Detail any special area of skill or expertise, for example, foreign body
analysis of food, cement, asbestos fibre characterisation, etc.
5. Indicate if you would be interested in special advertising facilities
linked to this feature.

The deadline for receipt of this information is 1st August, 1996. The
information will be compiled, tabulated, and published in the September
issue of Microscopy and Analysis. There is no charge to laboratories for
this service.

Information should be faxed to +44-1372-459957.


Dr. Larry Stoter
Technical Editor
Microscopy and Analysis






From: William Nicholson :      william-at-chem.gla.ac.uk
Date: Fri, 12 Apr 1996 11:32:19 +0100
Subject: Microscopy and Analysis Services in USA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe




From: BARBARA.HARTMAN-at-spcorp.com (BARBARA HARTMAN)
Date: Fri, 12 Apr 1996 08:52:46 -0400
Subject: Immunohistochemistry of Desmin and Actin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0

We have been utilizing immunohistochemistry to stain for smooth muscle
actin and desmin. We have noticed that actin stains all blood vessels
but desmin does not. Has anyone else noticed this ? Is there a
better stain to differentiate between smooth muscle and endothelial
cells ?




From: Dr. Peter Steele :      70152.3105-at-CompuServe.COM
Date: 12 Apr 96 10:35:06 EDT
Subject: Negative scanners TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have followed much of the thread on negative scanners in regard to TEM
negatives. It is still unclear if many biological/medical TEM labs use negative
scanners, and if so which make & model scanner has proven to be suitable.

Printing the digitized image obtained from the negative scanner is the other
half of the equation. What printers are in use for printing scanned negatives?
I realize that viewers of digitized negatives may have different expectations
than printing digitized images from other sources.

If you have experience with scanning TEM negatives (Kodak 4489) and printing,
especially from PCs (as opposed to MACs) I would appreciate hearing from you.

Thanks!





From: knecht-at-uconnvm.uconn.edu (David Knecht)
Date: Fri, 12 Apr 1996 09:15:20 -0500
Subject: Re: Stereo Imaging: Anaglyphs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I bought some red/blue glasses that were cheap and durable last year. I
had bought some others earlier that were crap as the lenses fell off
quickly.
The good ones were from:
Rainbow Symphony INc. 6860 Canby Ave. #120 Reseda CA 91335
800-821-5122

} At 10:39 AM 4/10/96 -0600, you wrote:
} } I am looking for red/green glasses for viewing some anaglyphs that I have
} } on slides. Ted Pella sells red/blue but I don't think they will work with
} } my pre-existing slides.
} }
} } Also, why the change from red/green to red/blue? Is this due to the "color
} } bombardment" phenomenon that causes a strain to some viewers or are there
} } other reasons?
} }
} } Any newer published references re anaglyphs besides the 1980 article by
} } Barber and Emerson in Scanning 3:202-206?
} }
}

Dr. David Knecht
Department of Molecular and Cell Biology
University of Connecticut
U-125
Storrs, CT 06269
Knecht-at-uconnvm.uconn.edu






From: Doug Cromey :      dcromey-at-ccit.arizona.edu
Date: Fri, 12 Apr 1996 08:20:20 -0700
Subject: histochemistry questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I apologize if you subscribe to both lists (confocal & microscopy) and
receive this twice. I wasn't sure where to send this one.

I have two histochemistry questions:

1) How does one block endogenous biotin (the tissue in question is insect)?

2) Are there "better" mounting media for sections that have had an alkaline
phosphatase reaction run on them (better than say, "permount")?

Thanks for your help.
Doug
.....................................................................
: Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy :
: Sr. Research Specialist University of Arizona :
: (office: AHSC 4212A) P.O. Box 245044 :
: (voice: 520-626-2824) Tucson, AZ 85724-5044 USA :
: (FAX: 520-626-2097) (email: dcromey-at-ccit.arizona.edu) :
:...................................................................:
http://www.pharm.arizona.edu/exp_path.html





From: Ulysses Garcia C. Lins :      ULINS-at-ibccf.biof.ufrj.br
Date: Fri, 12 Apr 1996 14:33:36 GMT-0300
Subject: Reprint request...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Sir,

I have read on the microscopy list, the suggestion of you paper on
Nanoplast. Please, if possible could you send me a reprint of this
article (Weyda F., 1990: Notes on the use of Nanoplast FB 101
for transmission electron microscopy. J.Electron Microsc.Tech., 16:
356-357), since I have been trying to embed magnetotactic bacteria in
Nanoplast with little success. Despite this, it is very difficult to
obtain a copy of this article, because of the absence of subscription
of J. Electron Microsc. Tech in my country.
Many thanks in advance,

Yours very truly,
Ulysses Lins

} Ulysses Lins
} Setor de Microscopia Eletronica
} Instituto de Microbiologia Prof. Paulo de Goes
} UFRJ - CCS - Cid. Universitaria
} 21949-900 - Rio de Janeiro - RJ
} Brazil
E-mail: imvuly-at-microbio.ufrj.br
you




From: Dr. Peter Steele :      70152.3105-at-CompuServe.COM
Date: Fri, 12 Apr 1996 15:10:16 -0500
Subject: Re: Negative scanners TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} I have followed much of the thread on negative scanners in regard to TEM
} negatives. It is still unclear if many biological/medical TEM labs use
} negative scanners, and if so which make & model scanner has proven to be
} suitable.
} Printing the digitized image obtained from the negative scanner is the other
} half of the equation. What printers are in use for printing scanned negatives?
*******************************************************************************
1. Scanning:

We use with great success for TEM negatives an AGFA Arcus II flatbed
scanner with the new AGFA FotoTune 2.0.8 Software. The negatives are
digitized to 14-bit with linear LUTs and 300 dpi for full frame images
(overview), and at optical 1,200 dpi for detail areas, each resulting in a
file size of 1Kx1Kx16-bit TIFF. The negatives should be exposed a step more
than used for printing to assure covered high lights. The gray tone quality
is extraordinarily good when compared to 8-bit scanning or cheap 8-bit
scanners, i.e., less than 1% IR (intensity range) noise, no scan lines even
at highest contrast resolution of contrasts smaller than three intensity
steps (from 16,000 steps per pixel), linear LUT (same contrast
distributions in highlights and shadows with positive or reversed negative
acquisition mode). We routinely zoom into the data 3-5 times with a "clean"
bicubic zoom in order to display details. The image quality is certainly
-although I cannot prove it at this time- better than can be established in
the photolab with silver halide printing.

2. Printing.

I tested several printers with a 1Kx1K image. Printer performances were
compared at the level of whole page images and 10x enlargements of the
printed images (LaserJet, Lazar Print, Dye sub,). Results are available at

http://panda.uchc.edu/htklaus/DigiLab/Printing.html

In short, you have to ask how many lines per inch are printed and at how
many gray levels (not always the same as dots per inch), plain paper or
special paper, permanence of the print, printing time, printing costs. I
print 300 lpi at 256 gray levels (4,800 dpi) on plain paper in 20-45 sec
for 1-10 MB of images per page at 3 cents per page with a HP LaserJet and
enhancer board in archival quality. I use the Lazar Print board. My source
was Smart Analytical Products (301) 598-8881. You will find everything at
my page. I am testing some more printers in the same manner (Epson, and a
1200 dpi Laser printer, but the results are expected not to be very
different from the already tested printers), A plain LaserJet will be a
good start for printing images.

Conclusion: Since over two years I have not used my state of the art
photolab since digital acquisition, flatbed scanning and digital LaserJet
printing fulfill all my needs for making image reports, slide templets
(photographed with 120 roll film on a copy stand), and publication prints.
However, my image communication is vastly improved through text imbedded
image printing in reports.

Best regards Klaus.

******************************************************************************
* : *
* Klaus-Ruediger Peters, Ph.D. : WWW Home Page: *
* Director, Molecular Imaging Laboratgory : *
* Biomolecular Structure Analysis Center : Molecular Imaging Laboratory *
* University of Connecticut Health Center : http://panda.uchc.edu/ *
* 263 Farmington Ave. : htklaus/index.html *
* Farmington, CT 06030-2017; U.S.A : Differential Hysteresis *
* : Processing Demo at http:// *
* Tel: (203) 679-3977; Fax: (203) 679-1989 : panda.uchc.edu./htbit/indiv/ *
* e-mail: Peters-at-BSAC.UCHC.EDU : /software_docs/dhp.html *
* : *
****************************************************************************
**






From: kaurin-at-rmslab.rockefeller.edu
Date: Fri, 12 Apr 1996 16:42:42 EST
Subject: immunogold/proteinase K

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have any experience using proteinase K to unmask
antigens on LRW sections? My use is for a cytoplasmic tail membrane label
using endothelial cell monolayers. What would you use to quench it?
Thanks for any suggestions!
Shelley Landon Kaurin









From: Bob Palm x8120 :      palm-at-aeetes.re.anl.gov
Date: Fri, 12 Apr 1996 16:55:07 -0500
Subject: Used SEM with EDS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



If anyone has or knows about a used SEM with EDS please contact me.

Robert Palm
Argonne National Lab
708-252-8120




From: MelanieOwl-at-aol.com
Date: Fri, 12 Apr 1996 17:25:45 -0400
Subject: Microscopy of Wax Crystals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello All:
A fellow employee of mine in a different department had some interesting
questions, some of which I can answer, some of which I have no experience
with which to answer. I was hoping that some of you may have some knowledge
you could share on this topic.
These questions all relate to cloud point measurements of oils using
polarized light microscopy, with a video camera and a temperature controlled
stage. A drop of oil is placed on a slide with a cover slip and heated above
the cloud point, then cooled 0.1 deg C/minute. The cloud point is determined
as the point at which the first wax crystal forms.

Her questions were:
1) Is it more preferrable to use a polarizing scope with a full circle 360
degree rotating stage or will an X-Y scope with a polarizer and polarizing
filter work just as well?

2) What kind of magnification would be needed to see the crystals form? Would
40X be sufficient? Would a zoom binocular be preferrable to fixed lenses?

3) Would there be any advantage to using a color camera over a black and
white one? The oil is black, and contrast is the most necessary element to
detect the crystals, not a color change.

4) What type of thermal stage is required? Mettler Toledo has a $12000
system, but is such a system overkill for this purpose? Physitemp has a TS-4
thermal stage with a 90 degree bend metal bracket with cooling fluid
circulating to it. The slide fits on the flat part of the bracket. This
costs about $3500. Will this stage be capable of ramping the temperature 0.1
degree C/minute? Are there other companies which sell thermal stages?

I was also wondering if anyone knows of any references on this test method.
Thanks in advance for your help.

Regards,
Melanie Behrens
Beacon, NY
behrema -at- Texaco.com
914-838-7261





From: becks-at-sunynassau.edu (Steve Beck)
Date: Sat, 13 Apr 1996 19:23:48 -0500
Subject: Ventilation of Chemical Storage Cabinets

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,

I am seeking the assistance of anyone who can point me to any specific
references (OSHA, NIOSH) on the proper storage of chemicals. Although our
institution has made a significant investment in the purchase of approved
chemical storage cabinets for acids, corrosives, and flammables, some of
our technical staff are reluctant to move a small quantity of chemicals
from a small storage room to a large, well ventilated preparatory room.
Their concern and claim is that these storage cabinets must be vented to
the outside in some manner. The chemicals of greatest concern include some
concentrated acid and base solutions (HCl, NaOH) and organic solvents such
as acetone and petroleum ether, in four to six liter quantities each.

I have searched the OSHA web site and have located information as to the
storage of flammables. Cabinet dimensions (metal or wood) relative to the
storage of specific quantities of flammable agents are noted, however,
there is no mention of the need for external cabinet venting. I have been
unable to find any reference as to the specific storage of acids or
alkaline solutions.

In most of the labs I have encountered (including my own EM lab), chemicals
are stored in the approved cabinets *without* separate external
ventilation. I would appreciate any specific references (web sites,
publications, etc.) that you could provide on this matter since I have a
dismantled TEM that has been awaiting installation into the small storage
room that these chemicals occupy, for over six months!

Please respond directly to me and I will post a summary to the group.
Thanks in advance!




Stephen Beck
Bio-Imaging Center/Electron Microscopy
Department of Biology
Nassau Community College
Garden City, NY 11530
Voice Mail: (516) 572-7829





From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Mon, 15 Apr 1996 08:42:46 +0100 (BST)
Subject: Re: BSE and C-J Disease

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


With reference to the BSE and CJD letters. While there is of course a
very faint chance that any of us can catch the any of the rarest of
diseases in the same way that we can be struck by lightning, the chances
of getting CJ from eating beef are marginally less that winning the big
prize on the UK National Lottery. I am quite content to eat muscle beef
and I buy a lottery ticket each week. I think you Americans have long had
the right idea in the you never eat meat that has come from the "engine
room". Leaving aside the bio-medical and econo-agricultural aspects of
BSE there is of course a delicious UK political and EURO political aspect
to this sorry story

Patrick Echlin
Cambridge UKOn Wed, 10 Apr 1996, Jane A. Fagerland (847) 935-0104
wrote:

} Thanks for all the comments concerning my remarks about prion diseases.
} I stand corrected on my assertion that humans don't usually eat central
} nervous system tissue from sheep and cattle. Several folks from the UK
} enlightened me - sheep and calf brains are considered a delicacy in some
} areas of the UK and are also used as fillers in hamburger and sausage.
} Also, soups are made from animals parts that may still have CNS tissues
} attached to them.
}
} Someone also pointed out that it was initially thought that scrapie
} could not jump from sheep to cattle; thus, there's precedent for inter-
} animal transmission, regardless of what seems likely right now.
}
} For anyone wanting more information on BSE, there is now a BSE group.
} To subscribe, send an e-mail to {majordomo-at-info.aphis.usda.gov} . Leave
} the subject line blank, and type {subscribe bse} in the message. There
} is also a BSE news and information service on the Web at
} http://www.cabi.org/
}
}
} Jane A. Fagerland, Ph.D.
} Dept. Microscopy and Microanalysis
} Abbott Laboratories
} Abbott Park IL 60064
}
}
}




From: SveEn-at-pai.liu.se (Sverker Enestrom)
Date: Mon, 15 Apr 1996 11:00:15 +0200
Subject: Re: histochemistry questions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




} I apologize if you subscribe to both lists (confocal & microscopy) and
} receive this twice. I wasn't sure where to send this one.
}
} I have two histochemistry questions:
}
} 1) How does one block endogenous biotin (the tissue in question is insect)?
* You can try successive 20 min incubations of sections in 0.1% avidin
and 0.01% biotin*

} 2) Are there "better" mounting media for sections that have had an alkaline
} phosphatase reaction run on them (better than say, "permount")?
* Yes, coverslip with aqueous-based medium after counterstaining*
} Thanks for your help.
} Doug

Good luck!
Sverker

*********************************************************
Sverker Enestrom M.D., Ph.D.
Department of Pathology
University of Linkoping, Sweden
Phone: +46 13 22 15 20
Fax: +46 13 13 22 57
*********************************************************






From: DAVID_GANTT-at-GSVMS2.CC.GASOU.EDU (DAVID G. GANTT)
Date: Mon, 15 Apr 1996 07:30:55 -0400
Subject: B&W SLIDES OF SEM NEGS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01530500ad97e36858ba-at-[141.165.35.119]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I USE TO USE A POSITIVE/NEGATIVE 35MM SLIDE FLIM TO MAKE POSITIVES OF SEM
NEGATIVES - DOES ANYONE IF THIS TYPE OF FLIM STILL EXISTS? IF SO PLEASE
GIVE ME ITS NUMBER AND WHO MAKES IT.

THANK YOU

Dr. David G. Gantt Phone: 1-912-681-5964
Dept. of Biology Fax: 1-912-681-0845
Landrum Box 8042 e-mail: david_gantt-at-gsvms2.cc.gasou.edu
Georgia Southern University
Statesboro, Georgia 30460-8042








From: Mary A. Molter :      molter-at-post.its.mcw.edu
Date: Mon, 15 Apr 1996 08:37:48 -0500 (CDT)
Subject: Re: Immunohistochemistry of Desmin and Actin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Although I have worked with this antibody in Prostate without success,
Factor VIII(von Willebrand factor), may be something to try. We feel
that we didn't have success because of the lack of vascularity in the
prostate to be able to observe any kind of staining patter. I have
two references that talk about using this antibody. The first reference
using a fluorescent Ab and the second uses it in cell culture, but either
should be of some help to you.
1. Potzsch G et al Laboratory Investigation 63(6):841-852 1990

2. McGuire PG and Orkin RW Laboratory Investigation 57(1):94-105 1987.

Good luck!

Mary
Laboratory Technologist
Medical College of Wisconsin
Department of Urology
molter-at-post.its.mcw.edu


On Fri, 12 Apr 1996, BARBARA HARTMAN wrote:

} Date: Fri, 12 Apr 1996 08:52:46 -0400
} From: BARBARA HARTMAN {BARBARA.HARTMAN-at-spcorp.com}
} To: histonet-at-pathology.swmed.edu, microscopy-at-aaem.amc.anl.gov
} Subject: Immunohistochemistry of Desmin and Actin
}
} We have been utilizing immunohistochemistry to stain for smooth muscle
} actin and desmin. We have noticed that actin stains all blood vessels
} but desmin does not. Has anyone else noticed this ? Is there a
} better stain to differentiate between smooth muscle and endothelial
} cells ?
}




From: GeneXs-at-aol.com
Date: Mon, 15 Apr 1996 10:33:10 -0400
Subject: Re: Immunohistochemistry of Desmin and Actin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


the e-mail address of Dr. Sam Bowser of the Wadsworth Center in
Albany N.Y. If someone knows it, I 'd appreciate if you could mail it to me.
Sam, if your out there, please e-mail me at genexs-at-mhv.net or
genexs-at-aol.com. Thankx.
Cya,
Gene Santagada





From: kna101-at-utdallas.edu
Date: Mon, 15 Apr 1996 08:06:36 -0500 (CDT)
Subject: Re: LM: Glycol metacrylate embedding media

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Michel,

I have used JB-4 for regular histology of the mouse skull base and had
good luck with it for toluidine blue stain only. I use 15-20 ml
breakaway molds, 0.7 gm catalyst to 100 ml of soln. A, and 1.0 ml of
solution B added to the embedding mixture of 100ml soln. A/0.7 gm
catalyst. I put the molds into icewater to start the embedding
procedure. I then overfill the molds, place the specimens into the soln.,
cover the surface with parafilm-making sure to force all the air bubbles
out from under the parafilm, seal the edges of the parafilm to the molds
with a hot spatula, and place the tray of ice water/molds into a
refrigerator overnite. If heat is a problem for your or
your friend's immunohistochemisty, the operation stated above must be done
quickly, as the polymerization of JB-4 is exothermic. The next day, I
remove the bocks from the refrig. and let them stand overnite at room
tempurature before I remove the molds. You might need to let the blocks
stand for a few more days before you attempt to cut them also. I cut
mine on a ralf knife made from a glass microslide ( the cheaper the
better) and I cut them on a standard rotary microtome with a razorblade
holder to hold the ralf knives.
As I stated above, I have not been able to get good staining with any
stain other than toluidine blue. If anyone out there has
suggestions on how to get any other stain to work with this material I'd
like to know.

Karen Pawlowski
On Fri, 12 Apr 1996, Michel Deschuyteneer wrote:

} Fellow microscopists,
}
} A colleague needs to embed nasal cavities from mice for immunocytochemistry.
} His antigens are fragile and do not whistand the standard embedding
} procedure with paraffin. Cryosectioning works but would preferably be
} avoided because of the bone tissue and decalcification has also proven
} detrimental to the antigenicity of his preparations.
}
} I suggested to try glycol metacrylate but I have no hands on experience with
} this material. In addition, there is a variety of commercial kits available,
} e.g. Polysciences' Immunobed or JB4, but the differences are not exactly
} obvious to me.
} Can anyone recommend a particular preparation of this medium? What are the
} pros and cons? Is there some good alternative?
}
} Any info and/or references would be welcomed.
} Thanks a bunch in advance.
} Regards,
}
} MICHEL
}
}
} ****************************************************
} Michel Deschuyteneer, Ph.D. deschuyt-at-sbbio.be
} Scientist Electron Microscopy Laboratory
}
} SmithKline Beecham Biologicals
} Rue de l'Institut, 89 B1330 Rixensart, BELGIUM
} Tel: +32-2-656 9290 Fax: +32-2-656 8164
} ****************************************************
} Disclaimer: the opinions expressed in this
} communication are my own and do not necessarily
} reflect those of SmithKline Beecham.
} ****************************************************
}
}




From: JOLOUGHLIN-at-MAT002.MATER.IE
Date: Mon, 15 Apr 1996 16:53:36 +0000 (GMT)
Subject: SUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






From: Jane A. Fagerland (847) 935-0104 :      FAGERLAND.JANE-at-igate.pprd.abbott.com
Date: Mon, 15 Apr 1996 11:51:00 -0600 (CST)
Subject: LM: staining of glycol methacrylate sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have a reliable method for hematoxylin and eosin staining of GMA
sections that works well for bone and soft tissue sections. I'd be
happy to mail or fax a copy of it to anyone who's interested.

Jane A. Fagerland, Ph.D.
Dept. Microscopy and Microanalysis
Abbott Laboratories
Abbott Park IL 60064






From: Larry Ackerman :      mishot-at-itsa.ucsf.edu
Date: Mon, 15 Apr 1996 12:21:45 -0700 (PDT)
Subject: B & W SLIDES

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In my opinion the best film for making positive slides from negatives is
Kodak Technical Pan Film 2415 (Cat No 129 7563 for 135-36 rollls
and 129 9916 for 150 ft rolls)developed in Kodak undiluted D-19 for 4
Minutes at 20 o C. You can also vary the developer and conditions in
order to adjust the contrast range--see Kodak Publication No. P-255. If
you want a positive slide from a positive image try Kodak Precision Line
film LPD4 (150ft rolls Cat No 157 8327) for a reasonable one step b & w
transparency.









From: rgwhite-at-vaxc.cc.monash.edu.au (Rosemary White)
Date: Tue, 16 Apr 1996 06:18:34 +1200
Subject: Re: B&W SLIDES OF SEM NEGS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Both Kodak and Ilford make a continuous tone direct positive B&W 35 mm
film. I have forgotten the specific product code, something like 6330 for
the Kodak film. It's apparently what they used to use for movies.

cheers,

Rosemary White
__ /
Department of Ecology _/ \__/ \
and Evolutionary Biology / \
Monash University / Australia \
Clayton, Victoria 3168 \ ____ /
phone 61-3-9905 5670 \_/ \_*_/
fax 61-3-9905 5613 __
email rgwhite-at-vaxc.cc.monash.edu.au \/






From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Mon, 15 Apr 1996 09:33:09 -1000 (HST)
Subject: Need suggestions for TEM immuno glycoproteins

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Aloha, microscopists,

I'm trying to help someone localize polysaccharides/glycoproteins in/on a
marine alga, first by using ConA or WGA, preferably labelled with
colloidal gold. Having never used ConA or WGA, I don't have a clue as
to where to begin. Can I use any glutaraldehyde? Left to my own devices
I will start with 0.25%. Paraformaldehyde? Like 1 to 3%? We will be
harvesting and fixing on Wednesday, April 17. After they're in blocks,
I can worry about the next steps. Blocking agents? Secondary
antibodies? If anyone has an opinion on how I should proceed, I would
love to hear from them!

Mahalo from out here in the middle of the Pacific...

Tina

*****************************************
Tina (Weatherby) Carvalho *
Biological Electron Microscope Facility *
University of Hawaii *
(808) 956-6251 *
tina-at-ahi.pbrc.hawaii.edu *
http://www.pbrc.hawaii.edu/bemf/ *
*****************************************





From: Donald Lovett :      lovett-at-trenton.edu
Date: Mon, 15 Apr 1996 17:53:48 -0400 (EDT)
Subject: Re: B&W SLIDES OF SEM NEGS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have had great success using Kodak Technical Pan film 2415 (develop at
maximum contrast, 4 minutes in D-19) directly through negatives sitting
on a light table, using 1/2 stop bracketing.

______________________________________________________________________
Donald L. Lovett e-mail: lovett-at-trenton.edu
Assoc. Professor, Dept. of Biology voice: (609) 771-2876
Trenton State College, NJ 08650-4700 fax: (609) 771-2674






From: tsi-at-werple.mira.net.au (Thomson Scientific:Paul Thomson)
Date: Tue, 16 Apr 1996 10:58:30 +1000 (EST)
Subject: Thomson Scientific Instruments Web Site

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {199604152322.SAA29765-at-IndyNet.indy.net}
To: Microscopy list {microscopy-at-Sparc5.Microscopy.Com}

Hello All,
For anyone who might be interested Thomson
Scientific Instruments now have a preliminary Web Site at the following URL:

http://werple.net.au/~tsi/

Please feel free to contact us for inquiries and support regarding WinEDS
and other products.


Regards and Thanks,



Paul Thomson
Technical Director
Thomson Scientific Instruments
http://werple.net.au/~tsi/






From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Mon, 15 Apr 1996 16:08:42 -0400 (EDT)
Subject: Re: B&W SLIDES OF SEM NEGS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Mon, 15 Apr 1996, Robert Schmitz, Biology wrote:

}
} } I USE TO USE A POSITIVE/NEGATIVE 35MM SLIDE FLIM TO MAKE POSITIVES OF SEM
} } NEGATIVES - DOES ANYONE IF THIS TYPE OF FLIM STILL EXISTS? IF SO PLEASE
} } GIVE ME ITS NUMBER AND WHO MAKES IT.
} }
} } THANK YOU
} }
} } Dr. David G. Gantt
} }
} Kodak #5302 Positive Release 35mm Film
} rschmitz-at-uwspmail.uwsp.edu
} or
} rschmitz-at-macsrv1.uwsp.edu
} (note its macsrv"one" not "el")
} Robert (Bob) J. Schmitz
} Department of Biology,
} University of Wisc. Stevens Point.
} Stevens Point, Wisconsin 54481
} ph 715-346-2420
}
David:
I used to use Tech Pan rated at ASA 125 and got beautiful slides.
Used a Nikon on automatic mode with an aperture of about f:6 on a 90mm lens.
Use a good light box.

Developed in D-19, 1:1 for 3min, fixed, washed and dried. If you use
Orbit Bath in your fix your fix time is cut to 2 min and wash 5 min.
Orbit Bath is good for printing also. Same times for prints.
For TEM slides did the same. In the past I used type 4489 EM film for my
slide. Tech Pan works rather well for making B&W slides from Negs.

Hope this helps.

Peace,
Phil Rutledge




From: Paul Webster :      paul.webster-at-yale.edu
Date: 15 Apr 1996 17:23:13 -0400
Subject: Re: immunogold/proteinase K

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1382533119.3257-at-QuickMail.Yale.edu}

Shelley Landon Kaurin writes:
" Does anyone have any experience using proteinase K to unmask
antigens on LRW sections? My use is for a cytoplasmic tail membrane label
using endothelial cell monolayers. What would you use to quench it?
Thanks for any suggestions! "

I presume that the antibody currently does not label the antigen in the LR
White. As you correctly guess, this is probably because the antibodies cannot
gain access to the antigen. However, I an not sure that treating the sections
with proteinase K will help expose the antigenic sites for you.

Firstly, the proteinase K is a protease and will not digest resins.

Secondly, if it did digest resin, how would you make sure that the proteinase K
did not digest your antigen during the incubation.

May I suggest that a better approach may be to permiabilize the cells prior to
embedding in resin, and then treating the sections with antibody, omitting any
resin digestion. The permiabilization will wash out much of the cytoplasm but
leave membrane proteins in place. Cytoplasmic tails are then accessible to
antibodies (unless they are really short and close to the membrane).

The permiabilization can be done either before or after fixation and can be as
simple as incubating living cells in hypotonic buffer, or incubating fixed cells
in detergent.

Need more? Contact me.

Paul Webster,
Center for Cell Imaging
Yale University School of Medicine






From: rgwhite-at-vaxc.cc.monash.edu.au (Rosemary White)
Date: Tue, 16 Apr 1996 09:21:41 +1200
Subject: Re: LM: Glycol metacrylate embedding media

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


RE:

} } A colleague needs to embed nasal cavities from mice for immunocytochemistry.
} } His antigens are fragile and do not whistand the standard embedding
} } procedure with paraffin. Cryosectioning works but would preferably be
} } avoided because of the bone tissue and decalcification has also proven
} } detrimental to the antigenicity of his preparations.
} }
} } I suggested to try glycol metacrylate but I have no hands on experience with
} } this material. In addition, there is a variety of commercial kits available,
} } e.g. Polysciences' Immunobed or JB4, but the differences are not exactly
} } obvious to me.

} } Regards,
} }
} } MICHEL


I haven't tried JB4 or glycol methacrylate for immuno work, but have used
butyl methyl methacrylate - it works well for plant tissue - for localising
microtubules, actin and callose at least. You need to add 5-10 mM DTT to
the fixative and to the embedding resin, seems to preserve the
antigenicity. The reference is Baskin et al. 1992 Planta 187: 405-413
(though the procedure we use is modified from this), in which earlier work
is cited also. Best results are with UV polymerisation in the cold, though
some people get away with room temperature polymerisation.

good luck,

Rosemary White
__ /
Department of Ecology _/ \__/ \
and Evolutionary Biology / \
Monash University / Australia \
Clayton, Victoria 3168 \ ____ /
phone 61-3-9905 5670 \_/ \_*_/
fax 61-3-9905 5613 __
email rgwhite-at-vaxc.cc.monash.edu.au \/






From: dianavd-at-eye.usyd.edu.au (Diana van Driel)
Date: Tue, 16 Apr 1996 12:22:21 +1100
Subject: Re: Osmium tetroxide / IGSS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: diana-at-pc-0.eye.usyd.edu.au
Message-Id: {v01540b01ad989eb333b2-at-[129.78.203.31]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Hello Everybody:
}
} I've received several comments recently about the fate of silver-enhanced
} gold particles (in EM) when specimens are treated with osmium tetroxide.
} The literature indicates that this process can cause a slight reduction in
} the size of the particles, probably due to re-oxidation of the deposited
} silver. However, the effect seems to be very variable: in some cases many
} of the particles disappear, or the color fades from the sections. Has
} anyone had problems with this, and has anyone found a good way to prevent
} it?
}
} Thanks,
}
} Rick Powell


Have just come back from holiday and it seems noone has answered this yet.
I do a lot of Ag-Au for EM and initially had just this problem. After
normal embedding of the Ag enhanced tissue, there would sometimes be
beautiful Ag particles, sometimes remains of Ag with holes where the
particles had been, sometimes nothing at all. The problem turned out to be
the uranyl acetate treatment. It appears (I am no chemist) that UAc
dissolves out the Ag in some way. Not using UAc cured the problem, but
obviously contrast was then hopeless. I now gold tone tissue - the Au-Ag-Au
complex is completely stable in osmium and UAc. There is an increase in the
contrast of the tissue, which may cause some people problems, especially if
high resolution is needed, but the method really works.

Gold toning: (after method of R. Arai et al 1992, Brain Res Bull 28:343-345)
after Ag enhancing, wash in water
0.05% gold chloride 10 mins/4deg
water wash
0.05% oxalic acid 2 min
water wash
1% sodium thiosulphate (freshly made) 1 hour
water wash and embed normally

Diana van Driel
Dept Ophthalmology
Sydney University
AUSTRALIA 2006






From: Richard Edward Bonshek :      RBONSHEK-at-fs2.scg.man.ac.uk
Date: Thu, 9 May 1996 02:39:53 GMT+1
Subject: Re: Need suggestions for TEM immuno glycoproteins

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Tina Carvalho {tina-at-pbrc.hawaii.edu}

} Date sent: Mon, 15 Apr 1996 09:33:09 -1000 (HST)
} From: Tina Carvalho {tina-at-pbrc.hawaii.edu}
} To: Microscopy Newsgroup {Microscopy-at-Sparc5.Microscopy.Com}
} Subject: Need suggestions for TEM immuno glycoproteins

} Aloha, microscopists,
}
} I'm trying to help someone localize polysaccharides/glycoproteins in/on a
} marine alga, first by using ConA or WGA, preferably labelled with
} colloidal gold. Having never used ConA or WGA, I don't have a clue as
} to where to begin. Can I use any glutaraldehyde? Left to my own devices
} I will start with 0.25%. Paraformaldehyde? Like 1 to 3%? We will be
} harvesting and fixing on Wednesday, April 17. After they're in blocks,
} I can worry about the next steps. Blocking agents? Secondary
} antibodies? If anyone has an opinion on how I should proceed, I would
} love to hear from them!
}
} Mahalo from out here in the middle of the Pacific...
}
} Tina
}
} *****************************************
} Tina (Weatherby) Carvalho *
} Biological Electron Microscope Facility *
} University of Hawaii *
} (808) 956-6251 *
} tina-at-ahi.pbrc.hawaii.edu *
} http://www.pbrc.hawaii.edu/bemf/ *
} *****************************************
}
---------------------------------------------------------------------

Hi,

Aldehyde fixation, even with gluteraldehyde, should not prevent
lectin binding. We've had good success with a wide range of lectins
on formaldehyde and glut. fixed human material, both LR white and
araldite embedded (we prefer LR white). To visualise bound lectin, we
used biotinylated lectin followed by gold labelled antibiotin. A
detailed method is given in:

Localisation of alpha(2,3) and alpha(2,6) linked terminal sialic acid
groups in human trabecular meshwork. SA Chapman, RE Bonshek, RW
Stodart, KR Mackenzie, D McLeod. British Journal of Ophthalmology,
1994; 78:632-637. }

We modified a method described by Slot and Geuze:

JW Slot and HJ Geuze. In: Polak JM and Varndell IM, eds.
Immunolabelling for electron microscopy. Amsterdam: Elsevier
Scientific. 1984: 129-142.

We've also visualised lectin binding (including ConA and WGA) in
resin embedded tisue at the LM level:

Glycoconjugates of the human trabecular meshwork: a lectin
histochemical study. SA Chapman, RE Bonshek, RW Stoddart, CJP Jones,
KR Mackenzie, E O'Donoghue, D McLeod. Histochemical Journal, 1995;
27:869-881.

This gives specificities, controls, etc, for a wide range of lectins.

Good luck!


Richard.









Richard Bonshek
Lecturer/Honorary Consultant in Ophthalmic Pathology
Department of Pathological Sciences
University of Manchester
Oxford Road
Manchester
M13 9PT
UK

TELEPHONE 44-161-276-5568
FAX 44-161-273-6354

E-MAIL Richard.Bonshek-at-man.ac.uk




From: rnessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Tue, 16 Apr 1996 07:57:14 -0500 (CDT)
Subject: LM stain cells prior to embeddment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi,
The question has been asked of me, what stain (prior to embeddment) to
use to be able to see a cluster of cells embedded in LR white. Since post
sectioning colloidal gold labeling is to be performed, the cells have not been
osmicated. They are translucent in the polymerized block. From my
understanding, the targeted gold binding site is a glycoprotein.
Does anyone have any suggestions as to what stain to use to be able to
visualize the cells, yet not interfer with labeling?

Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu





From: ubira-at-iqm.unicamp.br (Ubirajara Pereira Rodrigues Filho)
Date: Tue, 16 Apr 1996 08:50:54 -0200
Subject: SEM: TiO2 Grain size measurement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi;

Please can someone help me? I'm trying to measure the grain size of
TiO2 particles by SEM using a deposition of this particles from a 1% Renex
300 solution. Unhappiness I fail to separate the particles that are forming
agglomerates in this solution. Can someone give me some hint?

Ubirajara Pereira Rodrigues-Filho
Instituto de Quimica - UNICAMP
Campinas, SP, Brazil
e-mail: ubira-at-iqm.unicamp.br





From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 16 Apr 1996 12:30:20 -0400
Subject: HowToPolish Mo

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subject: Time: 7:25 AM
OFFICE MEMO HowToPolish Mo Date: 4/16/96

There is a graduate student in our department who has a single crystal rod of
Molybdenum. He wants to cut off thin slices, polish them so they are of
known orientation, flat, and free of surface cold-work, so he can use them
for some surface nucleation studies.

Is there anyone out there than could recommend a method for doing this that
might be accomplished in a reasonable time and with the equipment normally
found in metallography laboratories? If so, it would help him out a great
deal. Information on good, easily-controlled etchants and polishing
reagents for Mo might also be helpful.

Thanks, W. C. Bigelow (bigelow-at-umich.edu)





From: Marianne Ekwall :      Marianne.Ekwall-at-ah.slu.se
Date: Tue, 16 Apr 1996 16:39:12 +0100
Subject: Re: LM: staining of glycol methacrylate sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604161438.QAA07474-at-pinus.slu.se}
X-Sender: mekwall-at-pinus.slu.se
X-Mailer: Windows Eudora Light Version 1.5.2
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
microscopy-at-Sparc5.Microscopy.Com



At 11.51 1996-04-15 -0600, Jane A. Fagerland (847) 935-0104 wrote:
} I have a reliable method for hematoxylin and eosin staining of GMA
} sections that works well for bone and soft tissue sections. I'd be
} happy to mail or fax a copy of it to anyone who's interested.
}
} Jane A. Fagerland, Ph.D.
} Dept. Microscopy and Microanalysis
} Abbott Laboratories
} Abbott Park IL 60064
}
Dear Dr. Fagerland I would be very glad if yuo will be so kind and
send me a mail about the staing method for GMA sections.

Thanks Marianne Ekwall
Swedish University of Agricultural Sciences
Faculty of Veterinary Medicine
Dept of Anatomy and Histology}
}
}





From: Fermin, Cesar :      fermin-at-TMC.Tulane.Edu
Date: 16 Apr 1996 10:10:17 -0500
Subject: ER/Golgi-histo

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1382472539.90533-at-msmail.tmc.tulane.edu}

I need staining ER and/OR Golgi (ER best) on cultured cells with a
histochemical marker that would make the lumen electron dense. Some years
ago, I accomplished this (without wanting it) with a potassium chromate stain,
but can not remember the exact reference. If handy on somebody desk please
send a note directly to me. Gracias.

*********************************************************************
*Cesar D. Fermin, Ph.D \|T|/ Fax (504) 587-7389 *
*Tulane Medical School /|M|\ Voice Mail (504) 584-2618 *
*Pathology/SL79 \|C|/ Secretary (504) 584-2436 *
*New Orleans La 70112-2699 /|*|\ Lab/Techn. (504) 584-2521 *
* {Fermin-at-tmc.tulane.edu} ----} Prof. Pathology & Otolaryngology *
*http://www1.omi.tulane.edu/departments/pathology/fermin/cdftop.html*
*********************************************************************




From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Tue, 16 Apr 1996 13:55:55 -0500
Subject: Paraffin sectioning problem - liver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d00ad999cf057c8-at-[128.206.15.189]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We have a client who is fixing liver, brain, kidney and aorta with either
formalin or freshly depolymerized paraformaldehyde and embedding in
paraffin using our automated processor. Everything cuts fine except the
liver. Fixation time has run from perfusion only, perfusion + 4 hr,
perfusion + 24 hr, 4 or 24 hr immersion only. He describes the liver as
"turning to dust" as he sections it. All the other tissues (fixed and
processed at the same time) are cutting fine. Any liver histologists out
there with insight into this problem. Thanks.

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: South Bay Technology :      73531.1344-at-CompuServe.COM
Date: 16 Apr 96 14:00:00 EDT
Subject: Tripod Polisher Workshop Deadline

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



REDUCED FEE REGISTRATION DEADLINE APRIL 30, 1996

Workshop on Tripod Polishing

Workshop Objective
This course will cover all aspects of pre-thinning and focus on final thinning
via Tripod Polishing. Due to the limited class size and the extensive hands-on
opportuinities, this course is well suited to novices as well as advanced
Tripodders. The course will include sections on:

How to do it and why should I?
What's really going on and what am I really seeing?
How to prepare small, specific area cross-sections.
The problem of wildly differing materials (eg tungsten).
Rapid preparation of TEM cross-sections.
Preparation of a wide range of materials: semiconductors, ceramics, metals,...

Hands-on Opportunity
This course will be unique in that it will provide a hands-on opportunity for
every class participant. Tripod Polishers, Polishing Wheels, and pre-thinning
equipment will be made available to all participants and actual samples will be
prepared - by the students - as part of the course. This is a great opportunity
to get your hands dirty and actually learn by doing. The instructors will walk
you through each step of the process and then let you loose on the equipment.
This course is designed to teach the Tripod Polishing technique. Silicon
samples will be provided to the students and used as the basis for the course
teaching.

Workshop Location and Dates
South Bay Technology - San Clemente, CA
Dates: Friday & Saturday - June 14-15, 1996
Friday & Saturday - October 18-19, 1996

Previous Participants (partial list)
INTEL, AMD, Motorola, LSI Logic, Conner Peripherals, Univ of Maryland, Univ of
New Mexico, UNAM (Mexico), LG Electronics (Korea), Battelle, MEMC, MVA Inc.,
Univ of Michigan, U.S. Bureau of Mines, IBM, Naval Research Lab, Purdue Univ,
Univ of Alabama, Univ of Arizona, Univ of Colorado, Univ of Wisconsin.

Class Size
Due to the intensive hands-on aspects of this course, class size will be
strictly limited to 10 participants.

Registration Fee: $795 (includes lunches and Friday night Dinner)
$695 if registration fee paid by April 30, 1996 for June
Workshop and by August 31, 1996 for October Workshop

Registration Deadline: 30 days prior to workshop

For additional Information: Monica Pflaster
South Bay Technology, Inc.
1120 Via Callejon
San Clemente, CA 92673
TEL: 800-728-2233
FAX: 714-492-1499
e-mail: sbt-at-msa.microscopy.com




Registration Form

To register for the workshop, please fill out this form and send it, with
registration fee to:

South Bay Technology, Inc.
Workshop on Tripod Polishing
1120 Via Callejon
San Clemente, CA 92673

Payment must be made in the form of a check, money order, Visa or MasterCard.
Checks must be drawn on a U.S. Bank and made payable to South Bay Technology,
Inc. Credit card orders by FAX may be sent to South Bay Technology at
714-492-1499.

Name:


Affiliation:


Address:




City: State:
Zip: Country:_________
Telephone: FAX:

e-mail:________________________

Primary sample type:




VISA MasterCard Card #_________________________________

Expiration Date________ Signature of Cardholder_________________________

Cardholder name (Please print):________________________________________

Please circle workshop you are registering for: June 14-15 October
18-19





From: brannign-at-asrr.arsusda.gov (Peggy Brannigan)
Date: Tue, 16 Apr 1996 17:04:17 -0400
Subject: TEM- help with video camera/Mac

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello! Our EM lab has a DAGE-MTI SIT 68 video camera attached to the TEM
and I am looking for a way to get those images into a Macintosh 8500 (Power
PC) to analyze/process with the software IMAGE. The problem is that the
SIT 68 has an oddball line scan (2:1 interlace; 875/60) so I guess I'm
looking for a PCI board for the Mac that will grab and average the images.
Even though it's an AV Mac, it won't take this signal as is. Does anyone
have any suggestions, short of buying a new camera? Or having this one set
to standard? We'd like to keep the high resolution, if possible.

Also, once I finally have images on the Mac, I'd like to print them and
quit squandering time, money , environment etc in the darkroom. I've been
looking into 8 X 10 video printers but am wondering if anyone has had any
luck with some of the higher resolution laser printers that have come out?
I only need black and white.

Like most folks, we're pretty broke so I'm trying to do this on a
shoestring. Thanks for any help you can give me!

Peggy






From: ashamin-at-brauncorp.com
Date: Tue, 16 Apr 1996 17:46:18 -0500 (CDT)
Subject: E-mail on Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I've been receiving mail from your site, and I'd like to remove my name from
your list. I didn't sign up myself, but someone else may have. I'm not
even sure what your address is. If you want to talk to me about this, call
me at 612-683-8732. Thanks.





From: Robert414-at-aol.com
Date: Tue, 16 Apr 1996 20:48:31 -0400
Subject: KEVEX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a good INDEPENDENT KEVEX service engineer that services
the OH area? I am interested in having mapping installed on a KEVEX 8000
unit. If anyone has any suggestions please eMail me at robert414-at-aol.com.
Thanks.








Roberto
Garcia




From: oshel-at-ux1.cso.uiuc.edu (Philip Oshel)
Date: Tue, 16 Apr 1996 18:13:41 -0500
Subject: Re: LM stain cells prior to embeddment

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02120d02ad99d9e0d9f7-at-[128.174.23.164]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Hi,
} The question has been asked of me, what stain (prior to embeddment) to
} use to be able to see a cluster of cells embedded in LR white. Since post
} sectioning colloidal gold labeling is to be performed, the cells have not been
} osmicated. They are translucent in the polymerized block. From my
} understanding, the targeted gold binding site is a glycoprotein.
} Does anyone have any suggestions as to what stain to use to be able to
} visualize the cells, yet not interfer with labeling?
}
} Randy Nessler
} rnessler-at-emiris.iaf.uiowa.edu

Try 0.1% Neutral Red. May need to go stronger. In water, or buffer.
Phil

&&& Illigitimi non carborundum &&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&&
Philip Oshel
Center for Electron Microscopy
University of Illinois
(217)244-3145
oshel-at-ux1.cso.uiuc.edu

**** (Don't let the bastards wear you down) ***********************






From: John M. Libert :      jlibert-at-cpcug.org
Date: 4/16/96 8:01 AM
Subject: For Your Information - VIRUS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {317471CF.7C3A-at-cpcug.org}

The following message was passed on to me. I have no information as to the
veracity of this report or as to the ferocity of the virus, but take note.
Maybe others have more information.

Regards,
John Libert
Rockville, MD.


_____________________________ Forward Header
________________________________


} From Tom Beneifield, GA National Guard

______________________________ Forward Header
__________________________________


Joe,

This is serious stuff; thought you might like to review note I
received:



"VIRUS ALERT" "VIRUS ALERT"


1. This virus alert pertains to you if you have access to on-line
services such as AOL, CompuServe or any other service that allows you
to download files.

2. A new TROJAN HORSE virus has emerged on the internet with the name
PKZIP300.ZIP, so named as to give the impression that this file is a
new version of the PKZIP software used to ZIP(compress) files.

3. DO NOT DOWNLOAD THIS FILE UNDER ANY CIRCUMSTANCES!

4. If you install or expand this file, the virus will wipe your hard
disk clean and will affect modems at 14.4 baud rate and higher. This
is an extremely destructive virus and there is not yet a way to
cleaning up this destructive virus.

REPEAT: DO NOT DOWNLOAD ANY FILE WITH THE NAME "PKZIP300 reqardless of
the extension.

5. This message was confirmed by MicroSoft security.

CPT BREWTON
DOIM
""

See you,

Tom




From: Fermin, Cesar :      fermin-at-TMC.Tulane.Edu
Date: 16 Apr 1996 10:10:17 -0500
Subject: ER/Golgi-histo

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

by tmcpop.tmc.tulane.edu (8.6.12/8.6.12) with SMTP id LAA20505 for {Microscopy-at-
MSA.Microscopy.Com} ; Tue, 16 Apr 1996 11:01:00 -0500
Received: from tmcpop.tmc.tulane.edu (tmcpop.tmc.tulane.edu [129.81.174.79]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id LAA07688 for {Microscopy-at-spa
rc5.microscopy.com} ; Tue, 16 Apr 1996 11:01:17 -0500
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11) id LA
A07691 for dist-Microscopy; Tue, 16 Apr 1996 11:01:19 -0500
Received: from Sparc5.Microscopy.Com (sparc5.microscopy.com [206.69.208.10]) by
ormail.intel.com (8.7.4/8.7.3) with SMTP id SAA25166; Tue, 16 Apr 1996 18:39:24
-0700 (PDT)
Received: from ormail.intel.com by relay.hf.intel.com with smtp
(Smail3.1.28.1 #2) id m0u9MDN-000qDLC; Tue, 16 Apr 96 18:39 PDT




From: Y. Henry Sun :      mbyhsun-at-ccvax.sinica.edu.tw
Date: Wed, 17 Apr 1996 12:08:03 +0000
Subject: Zeiss CEM 902

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a Zeiss CEM 902 transmission EM. The LCD needs
replacement, and Zeiss is charging about US$4000. This seemed
outrageously expensive. Does anyone have experience like this?
Is this reasonable pricing? Any suggestions will be welcomed.
Please respond directly to mbyhsun-at-ccvax.sinica.edu.tw. Thanks
very much.

Henry




From: Jane A. Fagerland (847) 935-0104 :      FAGERLAND.JANE-at-igate.pprd.abbott.com
Date: Tue, 16 Apr 1996 16:26:00 -0600 (CST)
Subject: Hematoxylin and eosin staining of GMA sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta RANDB; Relayed; Tue, 16 Apr 1996 16:56:01 -0600
Mr-Received: by mta MCM$RAND; Relayed; Tue, 16 Apr 1996 16:56:12 -0600
Mr-Received: by mta RANDC; Relayed; Tue, 16 Apr 1996 16:56:24 -0600
Alternate-Recipient: prohibited
Disclose-Recipients: prohibited
Content-Return: prohibited

Since I received so many requests for this procedure, I decided it would
be faster to just send it via e-mail. If you have specific questions,
e-mail or telephone me directly.

REAGENTS:
Shandon Instant Hematoxylin:
available from Shandon/Lipshaw (1-800-547-7429), catalog number
9990107. Prepare according to package instructions.

Eosin/Phloxine B:
1% aqueous eosin 50.0 ml
1% aqueous phloxine B 5.0 ml
95% ethanol 390.0 ml
glacial acetic acid 2.5 ml

Scott's Solution:
tap water 1 liter
magnesium sulfate (anhydrous) 10 g
sodium bicarbonate 2 g


PROCEDURE:

1. Filter hematoxylin solution through Whatman No. 1 filter paper.
2. Stain sections in hematoxylin - 45 minutes
3. Rinse in running tap water - 5 minutes
4. Scott's solution - 2 minutes
5. Running tap water - 5 minutes
6. 70% ethanol - 10 dips
7. Eosin/phloxine B - 20 minutes
8. 80% ethanol - 10 quick dips
9. absolute ethanol - 2 or 3 changes, 5 dips each
10. xylene - 2 changes, 5 dips each
11. mount in Permount type medium and coverslip


I've used this for bone, lung, lymph node, liver and kidney. You can
vary the times in the stain solutions, depending on your preferences.
Intensity of eosin can be controlled by changing the time in 80%
ethanol.

Good luck!

Jane A. Fagerland, Ph.D.
Dept. of Microscopy and Microanalysis
D45M/AP31
Abbott Park IL 60064






From: Brett W. Cockman :      bcockman-at-uniwa.uwa.edu.au
Date: Wed, 17 Apr 1996 13:26:16 +0700 (WAST)
Subject: Water based mountants

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Hello all,

Could someone suggest a reliable (permanent) water - based mountant. I wish
to make wholemounts of muscle-nerve preparations which have been stored in
glycerin. Any help would be appreciated.

Brett



----------------------------------------------------------------------------
Brett W. Cockman
Technologist in Charge
School of Dentistry
University of Western Australia
Voice: (619-2205834)
Fax: (619-2213829)
e-mail; bcockman-at-uniwa.uwa.edu.au




From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Wed, 17 Apr 1996 01:10:43 EDT
Subject: SEM Sample Prep/TiO2 particles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

Ubirajara Iereira Rodrigues-Filho wrote the following:

Please can someone help me? I'm trying to measure the grain
size of TiO2 particles by SEM using a deposition of this particles from
a 1% Renex 300 solution. Unhappiness I fail to separate the particles
that are forming agglomerates in this solution. Can someone give me
some hint?



I am assuming you are talking about particle size of the TiO2 particles
and that the problem is in getting them to disperse. Let me describe
the method we have been using for more than twenty five years. It is
of unknown origin but it was taught to me by a very creative and
innovative microscopist many years ago who spent his career working at
the DuPont Experimental Station (Robert P. Schatz).

The "secret" is the magic mixture of 60% camphor/40% naphthalene which
when melted together, the two organic compounds form a eutectic which
itself melts at just a few degrees above room temperature.

The procedure is to make some of the liquid of the right composition,
e.g. 60%/40%, and then add on the order of 0.1-.0.5 wt% of the colloid
to be dispersed, in this case, TiO2. Stubborn behavior on the part of
the particles can be dealt with by way of a few minutes in an
ultrasonic shaker.

With any simple eye dropper, a drop of the liquid (containing the
particle dispersion) is put on the surface to be used as the substrate,
and since it is at room temperature, the liquid is almost
instantaneously frozen, resulting in a thin film coating on the
surface. If an SEM mount, then "surface" is the surface of the SEM
mount. If TEM studies are contemplated, a carbon coated glass slide
would be the surface. In either case, the end result is a substrate
surface covered with a thin frozen film of the eutectic system.

The interesting characteristic of the eutectic is that at room
temperature, in a vacuum evaporator using ordinary rotary pump vacuum
levels, the solid layer will sublime away (typically an over night
procedure), leaving the colloidal particles randomly scattered around
on the substrate surface. From this point, the sample can be prepared
either for SEM or TEM, depending on particle size. It has been our own
experience that for really careful studies on TiO2 particles, TEM will
give much better images from which measurements could be taken.

With regard to the camphor and naphthalene, nothing special is needed,
just ordinary off-the-shelf material of that type that is found in most
chemistry laboratory storerooms.

We have found this technique to be useful in the dispersion of a wide
range of inorganic particles or any other particle that would not be
effected by the camphor/naphthalene mixture.

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Wed, 17 Apr 1996 01:10:31 EDT
Subject: LM Glycol Methacrylate GMA resins

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

On Fri, 12 Apr 1996, Michel Deschuyteneer wrote:

} Fellow microscopists,
}
} A colleague needs to embed nasal cavities from mice for
immunocytochemistry.
} His antigens are fragile and do not whistand the standard embedding
} procedure with paraffin. Cryosectioning works but would preferably be
} avoided because of the bone tissue and decalcification has also
proven
} detrimental to the antigenicity of his preparations.
}
} I suggested to try glycol metacrylate but I have no hands on
experience with
} this material. In addition, there is a variety of commercial kits
available,
} e.g. Polysciences' Immunobed or JB4, but the differences are not
exactly
} obvious to me.

} Can anyone recommend a particular preparation of this medium? What
are the
} pros and cons? Is there some good alternative?

These kinds of questions are addressed in the following two
publications:

Stirling, John W., Histochemical Journal 24, 190-206 (1992)

Gerrits, Peter O., Eppinger, Bernhard, van Good, Harry, and Horobin,
Richard W., Cells & Materials, 1, No.3, 189-198) 1991

There can be some major differences in different GMA or HMPA based kits
coming from different sources. The first of the two above papers gives
specific mention about "low acid" GMA. The exact acid level can depend
on the purification process being used, the quality of the packaging,
and of course age of the product and the conditions of storage. The
acid level can make a big difference in the quality of the final
results. For LM, the acid component shows up as much higher levels of
background staining.

Standing in the SPI Supplies exhibit booth at trade shows might not be
the most perfect of scientific methods to draw conclusions, but the
impression I get is that because GMA has a viscosity like water, when
it comes to "bone tissue", it infiltrates quite nicely. Some have told
me "better than anything else". Another advantage of GMA is that
absolutely no alcohol dehydration is needed, since the monomer acts as
its own dehydrator. This is the presumed explanation for there being a
generally greater retention of antigenicity than with other commonly
used resins. The biggest perceived negatives associated with the use
of GMA seem to be the longer than normal learning curve frequently
reported in terms of learning how to work with it the first time.
Toxicity is often times cited as a negative, however, relative to most
of the other acrylates used in microscopy, it surely is not any worse,
and some have told me it does not effect them as much. Cost is also a
factor since the GMA and HPMA kits are certainly more expensive than
some alternative resin systems.

Disclaimer: SPI Supplies is the original commercial supplier of low
acid GMA and HPMA kits for both light and electron microscopy and we
would love to see more people enjoying the benefits of low acid GMA and
low acid HPMA embedding.

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: Dr. A. J. Garratt-Reed :      t.g_reed-at-fs2.mt.umist.ac.uk
Date: Wed, 17 Apr 1996 13:49:09 +0000
Subject: Re: SEM Sample Prep/TiO2 particles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604171315.OAA14894-at-mailspool.liv.ac.uk}
Comments: Authenticated sender is {t.g_reed-at-fs2.mt.umist.ac.uk}
GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)

Referring to Chuck Garber's ingenious method for dispersing
particulates for microscopy, what are the safety/health issues (if
any) of pumping the camphor/naphthalene mixture into the atmosphere?

Tony Garratt-Reed






From: Sandra F. Zane :      sfzane-at-unccvm.uncc.edu
Date: Wed, 17 Apr 1996 10:37:21 -0500
Subject: hierarchy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {2.2.16.19960417153721.21ef4c84-at-email.uncc.edu}
X-Sender: sfzane-at-email.uncc.edu
X-Mailer: Windows Eudora Pro Version 2.2 (16)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi Everyone,
There are a couple of questions which I would like to ask those of
you who are associated with small EM labs in a university setting.
Those of you who are kind enough to respond, would you share also
the number of people served by your facility and the number of technicians
employed?

1. Do your labs have a director?

2. If so, is that director a 9 month employee or a 12 month employee?

3. Is the director a faculty member or a staff employee?

4. What are the directors responsibilities?

5. If your labs do not have a director, is there a technician who assumes a
managerial position?

6. What are the duties of the technician/manager?

I will be most appreciative for all responses.

Sandra Zane
Sandra F. Zane, EM Tech. sfzane-at-email.uncc.edu
Dept. of Biology, UNCC Ph.(704)547-4051
9201 University City Blvd. Fax (704)547-3128
Charlotte, NC 28223





From: Bede Willenbring :      Bede.Willenbring-at-hbfuller.com
Date: Wed, 17 Apr 1996 12:45:34 -0500
Subject: SEM - Sputter/Coater Evaluations

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am in the process of replacing a sputter coater used for our SEM samples.
I am seriously considering three units: 1) Desk II with carbon yarn
evaporation accessory from Denton; 2) EFFACoater and EFFA Mark II Carbon
Coater from Ernest F. Fullam Inc; 3) SPI Module Sputter/Carbon Coating
System from SPI Supplies.

I would like to hear from anyone who has used one or more of these systems.
What did you like best, hate worst, overall performance, reliability.... If
you've used more than one, how would you compare them?

I'm considering including a thickness monitor. If you have one, do you find
it useful or a waste of money?

If you would rather not have you response included in a summary, please so
indicate.

My thanks to all who take the time to respond.
------------------------------------------------------------------------------
Bede Willenbring
H.B. Fuller Company

1200 Wolters Blvd.
Vadnais Heights, MN 55110

Email:Bede.Willenbring-at-HBFuller.com
Phone: (612) 481-3470
Fax: (612) 481-3309
------------------------------------------------------------------------------





From: Michael OKeefe :      Michael_OKeefe-at-macmail.lbl.gov
Date: 17 Apr 1996 10:53:41 U
Subject: About the PKZIP virus

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1382383916.54300-at-macmail.lbl.gov}

Subject: Time:9:59 AM
OFFICE MEMO About the PKZIP virus Date:4/17/96

Computerized microscopists:

For more information about the PKZIP300 virus, see --

http://www.nha.com/ciac6165.html

Michael A. O'Keefe
U of C, LBNL, NCEM






From: dbd1-at-uclink4.berkeley.edu
Date: Wed, 17 Apr 1996 11:17:13 -0700 (PDT)
Subject: Alternative to H&E for GMA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To All Who Are Interested,

There is a quick alternative to H & E staining of GMA sections. This works
on sections that are 0.5 to 2 microns thick. The stain is called Lee's
methylene blue-Basic fuchsin Stain. This stain takes about 1 minute and
gives some real nice staining similar to H & E. If you want anymore info
about this, contact me at psic-at-uclink4.berkeley.edu






From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 17 Apr 1996 14:16:20 -0400
Subject: RE- safety & TiO2 method

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1382371703.92132-at-mse.engin.umich.edu}

Subject: Time: 9:11 AM
OFFICE MEMO RE: safety & TiO2 method Date: 4/17/96

Camphor and naphthalene are both materials that are commonly used as moth
balls, and as such have been left laying around all over the place in
closets, drawers and storage chests in many homes throught the world for many
years. Probably the actual level of danger from them is minimal, but if you
bring the matter up, OSHA can probably find a lot of restrictions to apply
(as in the case for shipping distilled water into a laboratory).

We had a dean once who used to tell us, "If you don't want to hear the
answer you know I'll have to give, then don't ask the question!"





From: jean ross :      jeanross-at-emiris.iaf.uiowa.edu
Date: Wed, 17 Apr 1996 07:57:49 -0500 (CDT)
Subject: Re: Paraffin sectioning problem - liver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I've gotten liver to section by facing off the block to expose the tissue
and then soaking the blocks in ice water for about an hour. If you want
to do serial sectioning, you may need to soak the block again as you cut
deeper into the block.

Hope this helps you.

Jean Ross
Central Microscopy Research Facility
Univ. of Iowa
85 EMRB
Iowa City IA 52242
(319)335-8142
Web site: http://www.uiowa.edu/~cemrf



On Tue, 16 Apr 1996, Tom Phillips wrote:

} We have a client who is fixing liver, brain, kidney and aorta with either
} formalin or freshly depolymerized paraformaldehyde and embedding in
} paraffin using our automated processor. Everything cuts fine except the
} liver. Fixation time has run from perfusion only, perfusion + 4 hr,
} perfusion + 24 hr, 4 or 24 hr immersion only. He describes the liver as
} "turning to dust" as he sections it. All the other tissues (fixed and
} processed at the same time) are cutting fine. Any liver histologists out
} there with insight into this problem. Thanks.
}
} Thomas E. Phillips, Ph.D.
} Associate Professor of Biological Sciences
} Director, Molecular Cytology Core Facility
} 3 Tucker Hall
} University of Missouri
} Columbia, MO 65211
} (314)-882-4712 (voice)
} (314)-882-0123 (fax)
}






From: sphicae-at-rockvax.rockefeller.edu (Eleana Sphicas)
Date: Wed, 17 Apr 1996 17:45:00 -0400 (EDT)
Subject: Re: Paraffin sectioning problem - liver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Jill Craig {jcraig-at-unbc.edu}







From: John M. Libert :      jlibert-at-cpcug.org
Date: 4/16/96 8:01 AM
Subject: For Your Information - VIRUS

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604171809.OAA02218-at-thomas.ge.com}


} From Tom Beneifield, GA National Guard

______________________________ Forward Header
__________________________________


Joe,

This is serious stuff; thought you might like to review note I
received:



"VIRUS ALERT" "VIRUS ALERT"


1. This virus alert pertains to you if you have access to on-line
services such as AOL, CompuServe or any other service that allows you
to download files.

2. A new TROJAN HORSE virus has emerged on the internet with the name
PKZIP300.ZIP, so named as to give the impression that this file is a
new version of the PKZIP software used to ZIP(compress) files.

3. DO NOT DOWNLOAD THIS FILE UNDER ANY CIRCUMSTANCES!

4. If you install or expand this file, the virus will wipe your hard
disk clean and will affect modems at 14.4 baud rate and higher. This
is an extremely destructive virus and there is not yet a way to
cleaning up this destructive virus.

REPEAT: DO NOT DOWNLOAD ANY FILE WITH THE NAME "PKZIP300 reqardless of
the extension.

5. This message was confirmed by MicroSoft security.

CPT BREWTON
DOIM
""

See you,

Tom
----- End of forwarded message -----




From: K1JIA-at-vaxa.stevens-tech.edu
Date: Wed, 17 Apr 1996 20:03:39 -0500 (EST)
Subject: UNSUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Unsubscribe




From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Wed, 17 Apr 1996 21:27:00 EDT
Subject: TiO2 preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

Tony Garratt-Reed asked the following question:

Referring to Chuck Garber's ingenious method for dispersing
particulates for microscopy, what are the safety/health issues (if any)
of pumping the camphor/naphthalene mixture into the atmosphere?


Glad you raised the issue, it was one I have not thought about for a
long time. Of course as Wil Bigelow pointed out, part of what we are
talking about is just moth balls. And since such a small quantity is
actually being sublimed, the total volume is just one drop from an eye-
dropper, whenever I have myself walked into the lab area when this was
being done, I have never even smelled the naphthalene. But to be on
the safe side, I guess good practice would suggest that pumps be vented
to the outside, right?

Of course the whole set up when preparing the eutectic should be done
under a vented hood as should any work involving organic chemicals.

But since nothing seems to be present at levels high enough to smell,
perhaps incorrectly, we have not worried about it further.

With regard to my "teacher" of the technique, Bob Schatz, who is no
longer with us, he used to say that "the best things in life are those
that come free, and the best techniques in the EM lab are those that
don't require expensive materials".
How right he was.

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================





From: Marianne Ekwall :      Marianne.Ekwall-at-ah.slu.se
Date: Thu, 18 Apr 1996 08:15:04 +0100
Subject: Re: Alternative to H&E for GMA

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604180352.UAA27956-at-polaris.humboldt.edu}

At 11.17 1996-04-17 -0700, dbd1-at-uclink4.berkeley.edu wrote:
} To All Who Are Interested,
}
} There is a quick alternative to H & E staining of GMA sections. This works
} on sections that are 0.5 to 2 microns thick. The stain is called Lee's
} methylene blue-Basic fuchsin Stain. This stain takes about 1 minute and
} gives some real nice staining similar to H & E. If you want anymore info
} about this, contact me at psic-at-uclink4.berkeley.edu
}
} Hallo
}
Iwould like to have more info about the staining.
} Marianne Ekwall





From: MESJASZ-at-NACDH4.NAC.AC.ZA
Date: Thu, 18 Apr 1996 11:48:39 +0200
Subject: Al finder grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear All,


I am looking for Al finder grids. Does anyone know where to buy them (names,
fax/phone numbers and possibly e-mail address would be most heplful)

Many thanks in advance

Jolanta Mesjasz




From: Dr. A. J. Garratt-Reed :      t.g_reed-at-fs2.mt.umist.ac.uk
Date: Thu, 18 Apr 1996 10:43:08 +0001
Subject: Re: Boron analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Comments: Authenticated sender is {t.g_reed-at-fs2.mt.umist.ac.uk}

Dear Mark

There are two issues regarding light element analysis, namely,
does the X-ray reach the detector, and is the system capable of
recording it.

Regarding the first issue, there is more to the arrival of x-rays at
the crystal than just the "thin window". What is the window made
of? A polymer window can be quite good at transmitting B x-rays,
because the boron energy is below that of the carbon absorption
edge, provided the window is not too thick. However, on older
detectors there is, I am sure, a thick layer of ice, which may be
microns thick. This will do a fine job of absorbing boron x-rays
(and any other low-energy ones, too). Modern detectors (from some
manufacturers, at least) are quite able to withstand a cycle to room
temperature and a pump-out, and at least one manufacturer fits a
small heater to the crystal to allow it to be warmed, to sublime the
ice. If you have a TN5500, though, your detector may be fairly old.

Regarding the recording of the x-rays, one major problem is
system noise, which also degrades the spectral resolution. All
manufacturers have made steady progress at reducing the noise with
the passing of the years. However, noise can also be seriously
compromised by installation peculiarities, or inadvertent ground
loops, etc. This could be especially the case on an older
installation where, perhaps, a succession of people have been
responsible for it since the original installation, so that the
special arrangements made to reduce the noise originally have been
forgotten. Some systems allow you to view the "zero" peak which
enables you to judge directly the system noise. I don't know if the
5500 does this - many systems cut the spectrum off a few channels
above zero.

Yet another problem arises because of the very low x-ray yield
from light elements. If you analysing for boron in the presence of
heavier elements, than the boron peak is superimposed on the
bremsstrahlung from the heavier material. I did a
"back-of-the-envelope" calculation once that said that in LaB6 the
boron concentration is about 4 times the minimum detactable
concentration!. This would not be such a problem if you are
analysing boron metal, but could still be an issue for boric acid.

In the SEM, of course, you will only get a tiny boron signal,
even though the counts in the rest of the spectrum may be high,
because of the very large absorption of the boron x-rays in the
sample. You can minimise the effect by operating the SEM at a very
low voltage ( you only need 600eV electrons to excite the boron), but
of course you will lose source brightness and electron intensity (or,
conversely, spatial resolution) by doing this.

Good luck!

Tony Garratt-Reed.






From: Naresh Shah :      naresh-at-service1.uky.edu
Date: Thu, 18 Apr 1996 08:23:48 -0400
Subject: Re: Boron analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {31763454.5AAE-at-pop.uky.edu}

MARK DARUS (216) 266-2895 wrote:
}
}
} Hello,
} My EDX is a TN-5500 and I'm trying to determine if it is sensitive,
} enough to detect Boron. I have some standards, and in a moment I'm going to
} set up some boric acid and look into it. Just placing it in the chamber and
} getting the counts up, plus switching it to thin window doesn't give me
} overwhelming success from an SPI metal standard that I have.
} I'm sure there is much more to it than that. Is there any advice that you may
} have to offer. My beam current settings are 1, 5, 10, 20 & 30 KeV.
}
} Mark Darus
}
} Darus-at-cle.dnet.ge.com


A word of caution! Boric acid has high vapor pressure and will
sublime under SEM vacuum conditions contaminating your column.
I would suggest some other compound to test.

--

Naresh Shah

University of Kentucky
Consortium for Fossil Fuel Liquefaction Science (CFFLS) and
Department of Chemical and Materials Engineering (CME)
533 South Limestone Street, Room 111
Lexington, KY 40506-0043
Phone: (606)257-5119, Office: (606)257-4027 FAX: (606)257-7215
e-mail: naresh-at-pop.uky.edu




From: GeneXs-at-aol.com
Date: Thu, 18 Apr 1996 09:54:23 -0400
Subject: Microwave TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604181334.JAA17785-at-ns.ge.com}

Greetings people:
Can anyone e-mail me, or post, their experiences and opinions of microwave
processing for TEM. I work in a diagnostic pathology lab. We primarily want
to speed-up embedding time; but your comments concerning any stage of
processing will be appreciated.

Gene Santagada
genexs-at-mhv.net
genexs-at-aol.com

Kalvin Electron Microscope Lab.
Lenox Hill Hospital






From: Linda Fox :      LFOX1-at-wpo.it.luc.edu
Date: Thu, 18 Apr 1996 09:59:45 -0500
Subject: dual filter for fluorescent microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s1761241.084-at-wpo.it.luc.edu}
X-Mailer: Novell GroupWise 4.1

Does anyone know if there is a dual filter available for viewing FITC
and TEXAS RED at the same time? We have a Leitz Orthoplan 2. A
student is interested in purchasing this filter (perhaps used ?), if
the cost is reasonable.
Thanks .....also thanks for all the help with the Au target for
sputter coating. Linda lfox1-at-wpo.it.luc.edu





From: BARBARA.HARTMAN-at-spcorp.com (BARBARA HARTMAN)
Date: Thu, 18 Apr 1996 09:22:52 -0400
Subject: Antigen Retrieval and Desmin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0

We have been using antigen retrieval with desmin immunostain on mice
tissue and are getting some nice staining but not quite what we
expected. Our literature search has not turned up anything specific
involving the antigen retrieval and desmin , does anyone know of any
references or had any experience with this scenario ?




From: Joe D Geller :      geller-at-world.std.com
Date: Thu, 18 Apr 1996 11:01:56 -0400 (EDT)
Subject: JAMP-30 Computer control system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


attn: CFILION-at-BRMVM1.VNET.IBM.COM

Your mail was returned as undeliverable.
My company manufactures replacement computer control systems for the
JAMP-30 Auger system, but not parking stages. That is available from JEOL.

You may contact Geller MicroAnalytical Laboratory
at sales-at-gellermicro.com.

Thank you,
Joe Geller




From: Joe D Geller :      geller-at-world.std.com
Date: Thu, 18 Apr 1996 11:01:56 -0400 (EDT)
Subject: JAMP-30 Computer control system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


attn: CFILION-at-BRMVM1.VNET.IBM.COM

Your mail was returned as undeliverable.
My company manufactures replacement computer control systems for the
JAMP-30 Auger system, but not parking stages. That is available from JEOL.

You may contact Geller MicroAnalytical Laboratory
at sales-at-gellermicro.com.

Thank you,
Joe Geller




From: Joseph P. Neilly 708-938-5024 :      NEILLY.JOSEPH-at-igate.pprd.abbott.com
Date: Thu, 18 Apr 1996 08:04:00 -0600 (CST)
Subject: RE: SEM - Sputter/Coater Evaluations

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta RANDB; Relayed; Thu, 18 Apr 1996 08:08:36 -0600
Mr-Received: by mta MCM$RAND; Relayed; Thu, 18 Apr 1996 08:09:11 -0600
Mr-Received: by mta RANDB; Relayed; Thu, 18 Apr 1996 08:09:38 -0600
Alternate-Recipient: prohibited
Disclose-Recipients: prohibited
Content-Return: prohibited

Bede,

We have used a Desk II for several years and it has performed extremely
well. The only maintainence we have done is replace worn out targets
and chnage the pump oil once a year.

Joe Neilly
Abbott Laboratories
Dept. of Microscopy and Microanalysis
North Chicago, IL 60064






From: kna101-at-utdallas.edu
Date: Thu, 18 Apr 1996 08:02:00 -0500 (CDT)
Subject: Re: Paraffin sectioning problem - liver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I was just reading a "How to" chapter about problems of this sort. The
text is A manual for Histologictechnicians, Third edition, by Ann Preece,
Little, Brown and Co. Pubs., 1972. They state the reasons for this
crumbling could be inadequate fixation or incomplete dehydration.
Sometimes the tissue in the block is just too hard. The suggest trying
to soften the tissue first by placing it in water with a little detergent
added (1/2 tsp. per 100 ml H2O) and let it soak for up to 3 hours. If
this doesn't work, you may have to deparafinize the block and refix and
dehydrate and reembedd the block. Depending on how the tissue was
originally fixed, i.e. cardiac perfusion, the liver just may have not
gotten completely exposed to the fixative. I haven't tried these tricks
myself yet so if you have any luck, let me know. Good luck.

Karen Pawlowski

On Tue, 16 Apr 1996, Tom Phillips wrote:

} We have a client who is fixing liver, brain, kidney and aorta with either
} formalin or freshly depolymerized paraformaldehyde and embedding in
} paraffin using our automated processor. Everything cuts fine except the
} liver. Fixation time has run from perfusion only, perfusion + 4 hr,
} perfusion + 24 hr, 4 or 24 hr immersion only. He describes the liver as
} "turning to dust" as he sections it. All the other tissues (fixed and
} processed at the same time) are cutting fine. Any liver histologists out
} there with insight into this problem. Thanks.
}
} Thomas E. Phillips, Ph.D.
} Associate Professor of Biological Sciences
} Director, Molecular Cytology Core Facility
} 3 Tucker Hall
} University of Missouri
} Columbia, MO 65211
} (314)-882-4712 (voice)
} (314)-882-0123 (fax)
}
}
}




From: Richard Beanland +44 1327 356363 :      richard.beanland-at-gecm.com
Date: Thu, 18 Apr 1996 16:29:00 +0000 (GMT)
Subject: Backscattered electron imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Disclose-Recipients: prohibited

Hello All,
I have recently strted trying to do some quantitative backscattered
imaging of lead zirconium titanate (PZT). I'm getting rather puzzled by the
results and wondered if anybody had any ideas...

I naively took some papers at face value which state that backscattered
intensity is directly proportional to the mean atomic number of the material.
However, I am getting about twice the backscattered signal from PZT that I
expect. The composition has been confirmed by microprobe analysis, so I guess
that complex compounds and/or oxides behave differently from single elements
and binary metals (?)
I also see very strong differences in contrast between a nanocrystalline
phase (pyrochlore) and a large grained phase (perovskite), although both are
_supposed_ to be of the same composition. Am I actually seeing composition
differences or does the grain size have a large effect on backscattered
intensity?
Both effects are present when using a TEM (accelerating voltage 20-100kV)
and a FEGSEM (accelerating voltage 0.5-25kV).

Any ideas, references or advice would be very welcome!


Many thanks in advance,


Richard Beanland,
GMMTL
Caswell,
Towcester,
Northants NN12 8EQ
UK.





From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Thu, 18 Apr 1996 15:36:45 -0500 (CDT)
Subject: resume post

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have been asked to post this for Gary L. Van Landuyt:

OBJECTIVE:

Senior analyst position wanted.

EXPERIENCE:

15 years SEM, EDS, and WDS, operation, analysis and maintanence.
Physics instructor at a community college.
Technical writing.

Most recently employed at Bureau of Mines, Rolla, MO and Minneapolis, MN.

EDUCATION:

M.S. In physics -at- Northern Illinois University.

PROFESSIONAL AFFILIATIONS:

Microscopy Society of America
Microbeam Analysis Society
Minnesota Microscopy Society

ADDRESS:

Gary L. Van Landuyt
2730 West 66th Street Apt. 21
Richfield MN 55423-1971

PHONE:

H: 612-869-7439
W: 612-647-9322







From: Gary Login :      glogin-at-bih.harvard.edu
Date: Thu, 18 Apr 1996 14:17:21 -0400
Subject: Microwave TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Gene: There are several published reviews reporting how
microwave-accelerated processing methods are incorporated into routine
histopathologic practice. A recommended review is:
1. Leong, A. S.-Y. Microwave fixation and rapid processing in a large throughput
histopathology laboratory. Pathol 23: 271-273, 1991.

Reviews of microwave Fixation for TEM and LM are reported in:
1. Login, G. R., and A. M. Dvorak. Methods of microwave fixation for
microscopy. A review of research and clinical applications: 1970-1992. Prog
Histochem Cytochem 27/4: 1-127, 1994.
2. Kok, L. P., and M. E. Boon. Microwaves for microscopy. J Microsc 158:
291-322, 1990.
3. Leong, A. S.-Y. Microwave technology for morphological studies. Cell Vision
1: 278-288, 1994.

A review of microwave-accelerated embedding is in:
1. Giammara, B. Microwave embedment for light and electron microscopy using
Epoxy resins, LR White, and other polymers. Scanning 15: 82-87, 1993.


Two books describing how to get started using microwave techniques:
1. Kok, L. P., and M. E. Boon. Microwave Cookbook for Microscopists. Leyden:
Coulomb Press, 1992.
2. Login, G. R., and A. M. Dvorak. The Microwave Toolbook. A Practical Guide
for Microscopists. Boston: Beth Israel Hospital, 1994.


Gary Login (my e-mail address is at the end of this message)

In message {960418095422_472715458-at-emout13.mail.aol.com} writes:
} Greetings people:
} Can anyone e-mail me, or post, their experiences and opinions of microwave
} processing for TEM. I work in a diagnostic pathology lab. We primarily want
} to speed-up embedding time; but your comments concerning any stage of
} processing will be appreciated.
}
} Gene Santagada
} genexs-at-mhv.net
} genexs-at-aol.com
}
} Kalvin Electron Microscope Lab.
} Lenox Hill Hospital
}


Gary R. Login, D.M.D., D.M.Sc.
Assistant Professor of Oral Pathology
Beth Israel Hospital
Department of Pathology
330 Brookline Avenue
Boston, Massachusetts, 02215

glogin-at- bih.harvard.edu
Telephone: 617-667-2034
Fax: 617-667-8676





From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 18 Apr 1996 12:56:27 -0400
Subject: Safety Regs & water

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1382290113.76-at-mse.engin.umich.edu}

Subject: Time: 7:21 AM
OFFICE MEMO Safety Regs & water Date: 4/18/96

Following my comments on the safety regulations and the use of champhor and
naphthalene, several people have asked if OSHA does indeed require an MSDS
for water, and apply the usual chemical safety regulations to it. I assure
you that they actually and in fact do. Here is a message on the subject I
received recently from one of my ex-students:
- - - - - - - - - - - - - - -
"At Lockheed, the chem lab had a high resolution chromotography system
requiring triple distilled water. The bottles had to be shipped with an
MSDS, they were labeled "Hazardous", and they had to be inventoried as a
"Chemical". Likewise, an NBS EDS standard of stainless steel had to be
handled as a "Hazardous Material" because it contained chromium!"
- - - - - - - - - - - - - -
Further support can be found in a column written by Mike Ryoko, a syndicated
columnist whose writings appear in many newspapers throughout the country, a
couple of days ago. In this column he referrs to a similar situation that
was called to his attention in the Chicago area. He of course makes a number
of witty comments about the situation.

There are, of course, many other examples of situations where safety
regulations have been enforced to a ridiculous end. I heard of one
laboratory where a bottle of sand, which was unfortunately had the chemical
formula SiO2 written on it, was declared a hazardous substance (presumeably
because it contained the element silicon).

We used to have a foundry in our department which used large quantities of
ordinary sand for various purposes. Again some of this material was in bags
labeled "Silica Sand", and when workers who were cleaning out the foundry
area spotted this they called in a full safety crew complete with gas masks,
sealed suits, and all related equipment, to dispose of the material. This
was done in the summer, and undoubtedly some of the workers spent their
weekends picnicking at the beach without giving the hazard of doing so a
second thought.

In a high school not far from here a student accidentally knocked over a
graduated cylinder, spilling about 100 ml of ethyl alcohol. The entire
school was closed down for the remainder of the day while the local fire
department was called in to clean up the spill. It was later learned that
several teachers celebrated the unexpected vacation by stopping at a local
bar and consuming more alcohol than was spilled at school.

These are only a few examples that I know of. There undoubtedly many more
floating around, and it would perhaps be interesting to hear about them.

I don't mean to imply that I am against appropriate safety regulations, and I
do recognize that the imposition of such regulations have been very
beneficial in protecting people in a great number of situations. However,
in some instances, largely because of the ignorance and perversity of those
who formulate and enforce these regulations, the system has been pushed to
ridiculous extremes.

In closing, I should correct myself. I said that camphor was commonly used
in moth balls. I think this is incorrect. The two common ingredients for
moth balls are, I believe, naphthalene and para-dichlorobenzene. I know that
naphthalene is still used in this way, because I recently bought some
naphthalene moth balls to evict a possum from a hole it had dug under my
front porch. However, the para-dichlorobenzene may not be so commonly used
in this way any more because it is a chlorinate phenyl compound. Camphor,
on the other hand, has been commonly used in the stuff you spread on your
lips to help heal chapping, and for years was used as a non-prescription
medicinal in such things a camphorated oil, nazal decongestants, cough
lozengers, etc.





From: GVKM07A-at-PRODIGY.COM (DR CHARLES A GARBER)
Date: Thu, 18 Apr 1996 20:07:58 EDT
Subject: Detection of Boron

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Charles A. Garber, Ph. D. * EMC.Ver #2.10P ] --

Mark Darus wrote:

My EDX is a TN-5500 and I'm trying to determine if it is sensitive,
enough to detect Boron. I have some standards, and in a moment I'm
going to set up some boric acid and look into it. Just placing it in
the chamber and getting the counts up, plus switching it to thin window
doesn't give me overwhelming success from an SPI metal standard that I
have. I'm sure there is much more to it than that. Is there any advice
that you may have to offer. My beam current settings are 1, 5, 10, 20
& 30 KeV.
==================

There is just no better "standard" than a solid lump of pure boron! It
is inert and it won't be changed by the electron beam. Boric acid will
be unstable and as others pointed out, it will sublime causing all
kinds of other problems.

Based on our own experience fielding technical service calls, you
should check to make sure no one has inadvertently applied too thick of
a carbon coating. People don't admit to doing this but it does happen!
This has in fact happened before, with some element of frequency, where
the carbon was thick enough to absorb all exiting B x-rays. If that is
indeed the case, then read the User Manual that came with the SPI
standard for instructions for cleaning it up and returning it to a
state where you will see the x-rays. Note: Great care must be taken
or you can run the risk of damaging the electron beam lettering
opposite each standard item.

Another point: You have listed "beam currents" but these are really
accelerating voltages. You did not mention the current, but if you see
any x-ray continuum at the B (K alpha) energies, then one can assume
the detector is working. To be sure about that, check to make sure you
can see the slightly more energetic C (K alpha) from the carbon
standard.

More often than not, the problem is simply that the detector, for one
reason or another is just not "up" to it. But you should check whether
it is a transient kind of thing (e.g. ice) or something more serious.

Additionally, and no one likes to think this way, what if something
horrible did happen and something did get mixed up along the way? But
"it" still does have to be "something" and the x-rays from that
something surely should be able to be measured and the material
actually present identified.

Hope this information might be helpful to you.

Chuck

======================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail:
GVKM07A-at-prodigy.com
West Chester, PA 19381-0656 USA Customer Service:
SpiSupp-at-aol.com

########################################################
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html
########################################################
======================================================








From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Fri, 19 Apr 1996 17:49:28 +1100
Subject: EM: Reichert OMU2 to give away

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: st004718-at-brandywine.otago.ac.nz
Message-Id: {v01540b02ad9cd5dbaaa0-at-[139.80.120.139]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


Dear Microscopists,
We have two Reichert OMU2 ultramicrotomes which we would like to give away
(either whole or whatever pieces people want) to anyone prepared to cover
freight costs.

They are mid 1960's vintage, one is complete and one is missing a few bits,
like the binoculars. They both still have their antivibration tables.
I know it is long shot, but they may be useful for someone out
there...otherwise to the tip they go.

Contact me if interested.

Yours faithfully,

Richard


Richard Easingwood
South Campus Electron Microscope Unit
School of Medical Sciences
University of Otago
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

SOUTHERNMOST E.M UNIT IN THE WORLD















From: Keith Ryan :      KPR-at-wpo.nerc.ac.uk
Date: Fri, 19 Apr 1996 08:41:09 +0000
Subject: Safety Regs & water -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s1775117.014-at-wpo.nerc.ac.uk}
X-Mailer: Novell GroupWise 4.1

Hello all

This is a slightly facetious input re. safety data sheets!

In the UK there is a very useful pair of publications from BDH, a chemicla
supply company (used to be known as British Drug House, I believe).
These are collcted data sheets. Yes folks, there is one for water. Here are
a few salient points which all users of the substance should bear in mind at
all times:

1. colourless liquid - maybe that implies it is rare and difficult to see if
dropped!
2. Against solubility in water: miscible in all proportions.
3. fire and explosion hazard: not applicable (thats a relief - although if it
caught fire I suppose one could foolishly attempt to extinguish with more
water?).
4. Health hazard: no significant hazard expected, may be irritating to the
eyes.
5. Toxicity: no data.
6. Carcinogenicity: no evidence of carcinogenic properties.
7. First aid - eyes: irrigate thoroughly with water(!)
8. First aid - lungs: remove from exposure.
9. First aid - skin: wash off thoroughly with soap and water (work that one
out!).
10. First aid - mouth: wash out thoroughly with water. In severe cases
obtain medical attention(!)
11.Reactive hazards: violent reaction with acyl halides, alkali and alkaline
earth metals ....
12. Spillage disposal: wear appropriate protective clothing - listed are
gloves, goggles, face shield and protective apron. Luckily, respirators are
not needed.

Have a nice day (but watch out for water raining down!).

Keith Ryan
(Local Safety Advisor - among other hats worn!)
(used to be a good EM guy - there's a cry from the heart!)
Plymouth Marine Laboratory
Citadel hill
Plymouth PL1 2PB,
England





From: AWBlackwoo-at-AOL.com
Date: Fri, 19 Apr 1996 09:53:44 -0400
Subject: Aluminum Finder Grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


19 April 1996

Jolanta Mesjasz asked about the availability of aluminum index or finder
grids.

As some of you know, I am responsible for the laboratories of Structure
Probe, Inc., parent company of SPI Supplies. We would love to be able to
sell finder grids and other specialty products in aluminum, but we have not
been able to develop products that meet our quality requirements. The
reason, unfortunately, is very simple but also very basic.

Most TEM grids are made by electrodeposition. The copper, nickel, gold and
other precious metal grids which we and our competitors offer are plated.
While there is a lot of technology involved in the process, it is basically
very simple, and it allows control over bar width and other details to an
incredible degree; letters, numbers, fiduciary marks and other
identifications are easy compared to controlling bar width within a range of
micrometers of the nominal dimension.

For aluminum, tungsten, molybdenum, beryllium and stainless steel grids,
however, the production process uses etching. There are several problems,
but they come down to problems in the resolution of the imaging process, the
thickness of the grid material and the nature of the etching process. We
tear our hair out trying to put out product which LOOKS like a grid, with
bars, etc. Details like letters and numbers are simply beyond our capability
at this point; we do not know of a process which can produce these details at
a price anyone is willing to pay.

If a reader knows how to produce an aluminum (or other etched) finder grid,
we (and I suspect our competitors, as well) would be most happy to talk with
you about setting up production.

Andy

Andrew W. Blackwood, Ph.D.
Structure Probe, Inc.
63 Unquowa Road
Fairfield, CT 06430-5015
Ph: 1 203 254 0000
FAX: 1 203 254 2262
e-mail: AWBlackwoo-at-aol.com
WWW: http://mail.cccbi.chester.pa.us/spi/spihome.html






From: peling-at-amnh.org (Peling Melville - Interdepartmental Facilities)
Date: Fri, 19 Apr 1996 09:43:41 -0500
Subject: Re: SEM - Sputter/Coater Evaluations

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510102ad9d53dd4f34-at-[198.116.4.5]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

On Wednesday April 17th, Bede Willenbring writes:

} I am in the process of replacing a sputter coater used for our SEM samples.
} I am seriously considering three units: 1) Desk II with carbon yarn
} evaporation accessory from Denton; 2) EFFACoater and EFFA Mark II Carbon
} Coater from Ernest F. Fullam Inc; 3) SPI Module Sputter/Carbon Coating
} System from SPI Supplies.
}
} I would like to hear from anyone who has used one or more of these systems.
} What did you like best, hate worst, overall performance, reliability.... If
} you've used more than one, how would you compare them?
}
} I'm considering including a thickness monitor. If you have one, do you find
} it useful or a waste of money?
}
} If you would rather not have you response included in a summary, please so
} indicate.
}
} My thanks to all who take the time to respond.


We purchased a Denton Desk II with the carbon yarn evaporation accessory
acouple of years ago and have had no problems with it. The only things
that I have had to do in terms of maintenance is give it one oil change a
year, target changes, and cleaning of the glass cylinders & various
removable stage parts.

Peling Melville


--------------------------------------------------------------
Peling Fong Melville
Senior Scientific Assistant
Interdepartmental Facilities
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 U.S.A.
******************************
E-mail: peling-at-amnh.org
Work #: (212) 769-5469
FAX #: (212) 769-5495






From: Schoonhoven, Robert :      rschoonh.sph-at-mhs.unc.edu
Date: Fri, 19 Apr 1996 11:59:50 -0400
Subject: Re: LM: Glycol metacrylate embedding media stainig procedures

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {3C80773101F70300-at-mhs.unc.edu}
In-Reply-To: {8E6A773101F70300}

In a past life Sandy Brady and I gave several workshops on GMA processing
and staining. If I remember we gave out a very comlete handout of
staining techniques. I don't know if I can easly locate a copy but there
should be some copies floating aroud somewhere. Perhaps Sandy has an
extra copy or two.
regards
bob

Robert Schoonhoven
Laboratory of Molecular Carcinogenesis and Mutagenesis
University of North Carolina
CB#7400, Rosenau Hall
Chapel Hill, NC 27599-7400
Ph. 919-966-6343
Fax 919-966-6123





From: Jill Craig :      jcraig-at-unbc.edu
Date: Sat, 20 Apr 1996 09:10:12 -0700 (PDT)
Subject: sodium sulfate in SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi,

I have been asked to image crystals of nearly pure sodium sulfate. As it
is a drying agent which contains 10 molecules of water per molecule of
Na2(SO)4 I'm a little concerned about putting it in the SEM. I have put
it in a drying oven for 2 days and I was planning to put the stub under
vacuum for a while before coating it with gold. Are there any other
precautions or preparations I should do?

Thanks for the help!

Jill




From: Schoonhoven, Robert :      rschoonh.sph-at-mhs.unc.edu
Date: Fri, 19 Apr 1996 13:24:23 -0400
Subject: Re: Paraffin sectioning problem - liver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {61B7773101F70300-at-mhs.unc.edu}
In-Reply-To: {FD69773101F70300}

I'm assuming that you are working with rat or mouse models. We do a lot
of rat, mouse and fish liver work in our lab. All of our histology
processing is contracted out so I had to develop SOP's for processing
and sectioning. If you want I can fax these out to you. We've had no
sectioning problems (over 5 years now) since these SOP's were written.
regards, bob

Robert Schoonhoven
Laboratory of Molecular Carcinogenesis and Mutagenesis
University of North Carolina
CB#7400, Rosenau Hall
Chapel Hill, NC 27599-7400
Ph. 919-966-6343
Fax 919-966-6123





From: rnessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Fri, 19 Apr 1996 12:43:13 -0500 (CDT)
Subject: "Free" Zeiss TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi All,
A department here at the University of Iowa contacted us about finding a
new home for their used Zeiss 10C. It has been under service contract until
just recently. Since they have decided not to renew the service contract, they
are interested in finding someone who is willing to pay for the system to be
dismantled, shipped, and reassembled. These are the only costs, as they will
give to the instrument to said party. They have pressing needs for the space
that the scope occupies, and thus are considering junking it. It would
almost be a shame to recycle a fully functional microscope......
If interested, contact Kenneth Moore (kenneth-moore-at-uiowa.edu) or myself
either by email or phone (319-335-8142).

Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu






From: GANTZ-at-med-biophd.bu.edu
Date: Fri, 19 Apr 1996 12:19:59 -0400 (EDT)
Subject: Fire Safety

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think that the fire safety issue is important enough to continue
the thread. An incident happened in my building last week which reminded
us of the importance of smoke detectors. An instrument which had been
left running and unattended many times in the past began to melt down
when on this occasion was left unattended for 15 minutes. Part of the
instrument became red hot, melted a hole in the bench top, and sent thick,
black, acrid smoke into the room and two adjacent ones. Directly above
the instrument are located air intake and exhaust vents which pulled
some smoke and presumably heat out of the room. The heat activated
sprinkler (designed to be activated at 135F, we think) did not activate,
presumably because heat was exhausted away. Fortunately, a student
returned to discover the situation.
Later we extracted the following information from the large
gathering which included 7 firemen, several safety and facilities
management personnel: smoke detectors are not required under current
fire safety codes in this city. Our building is two years old. The
nearest smoke detector was approx. 100 feet away in the lobby.
I suspect that the lack of smoke detectors in laboratories is
not unique to this building or this city. The safety office here on
campus has at least taken this matter under consideration. We also
realized that we have taken equipment reliability for granted.

Don Gantz
Boston Univ. Med. School




From: hallel :      hallel-at-macgw1.crd.ge.com
Date: 19 Apr 1996 15:18:09 U
Subject: Job Opening/GE CRD/EBSP-SEM Operator

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1382195461.81071-at-macgw1.crd.ge.com}

The Microscopy and Microanalysis Program in the Materials Characterization
Laboratory at GE Corporate Research and Development, Schenectady, NY, has an
opening for a Microanalyst/Microscopist at the Professional/Lead Professional
(BS/MS) level. The primary duties associated with this position involve the
execution of research projects involving Electron Backscatter Pattern (EBSP)
analysis of crystalline materials in the Scanning Electron Microscope (SEM).
Additional duties may involve research conducted using a variety of other
electron imaging and analysis equipment, including transmission electron
microscopes and electron microprobes. The Microscopy and Microanalysis
Program at GE CRD is one of the world's leading centers for the application of
EBSP techniques to the analysis of texture and orientation. Two dedicated
SEMs are used for this capability. The present position involves the
operation and maintenance of that equipment, and the management of project and
sample flow through the facility.

The Materials Characterization Laboratory is involved in research into the
structure and composition of materials in support of development programs both
at GE CRD and at GE businesses. A wide range of materials, including metals,
ceramics, composites, polymers, and coatings/surface modifications, are
analyzed by this group. Staff members are expected to work independently with
a high level of expertise, and to become involved with a number of major
project teams. Good communication skills, both written and oral, are
extremely important.

A BS or MS in Materials Science or a closely-related field is required. Prior
experience with electron beam instruments, particularly transmission and
scanning electron microscopes, is highly desirable. Some familiarity with
crystallography and computer usage/programming in the Windows environment is
also desired.

Resumes and other information can be sent to:

Ernest L. Hall
Manager, Microscopy and Microanalysis Program
Room K1-2C27
GE Corporate Research and Development
PO Box 8
Schenectady, NY 12301
Fax: 518-387-6972
E-mail: hallel-at-crd.ge.com





From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 19 Apr 1996 17:16:45 -0400
Subject: RE- PKZIP300 virus

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1382188116.34333-at-mse.engin.umich.edu}

Subject: Time: 12:15 PM
OFFICE MEMO RE: PKZIP300 virus Date: 4/19/96

Does anyone know whether the PKZIP300 virus reported by Scott Walck is
directed at Macintosh or IBM computers?? or can it infect both??
W. C. Bigelow (bigelow-at-umich.edu)





From: William Tivol :      tivol-at-wadsworth.org
Date: Fri, 19 Apr 1996 14:20:14 -0400 (EDT)
Subject: Re: Backscattered electron imaging

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Richard,

} Hello All,
} I have recently strted trying to do some quantitative backscattered
} imaging of lead zirconium titanate (PZT). I'm getting rather puzzled by the
} results and wondered if anybody had any ideas...
}
} I naively took some papers at face value which state that backscattered
} intensity is directly proportional to the mean atomic number of the material.
I just taught my radiation sciences class about nuclear backscatter-
ing (it was used as an analytical method on the moon, BTW), and the text--
Nuclear and Radiochemistry, Friedlander, et al., p430--states that the cross
section is proportional to Z^2. Since it is a Coulomb scattering, and since
e-e back scattering would have zero energy, I would expect electron back
scattering to have the same Z-dependence as proton backscattering. Per-
haps David Joy will comment.

} However, I am getting about twice the backscattered signal from PZT that I
} expect. The composition has been confirmed by microprobe analysis, so I guess
} that complex compounds and/or oxides behave differently from single elements
} and binary metals (?)

This shouldn't be the case for backscattering. There might be minor
differences for low-angle scattering due to differences in the valence elec-
tron distribution--the low-order CBED amplitudes are sensitive to this.

} I also see very strong differences in contrast between a nanocrystalline
} phase (pyrochlore) and a large grained phase (perovskite), although both are
} _supposed_ to be of the same composition. Am I actually seeing composition
} differences or does the grain size have a large effect on backscattered
} intensity?

Does the effect vary with the orientation of the large grains?

} Both effects are present when using a TEM (accelerating voltage 20-100kV)
} and a FEGSEM (accelerating voltage 0.5-25kV).

I am not surprised that you see no dependence on voltage.
Yours,
Bill Tivol




From: hallel :      hallel-at-macgw1.crd.ge.com
Date: 19 Apr 1996 15:44:31 U
Subject: Call for Papers/Fall 1996 MRS/Interface Engineering

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {n1382193754.78269-at-macgw1.crd.ge.com}
"Carter, Barry" {carter-at-cems.umn.edu}
X-Mailer: Mail*Link SMTP-MS 3.0.2

Call for Papers

Symposium W: Interfacial Engineering for Optimized Properties
Fall 1996 Materials Research Society Meeting
December 2-6, 1996

Description:
Interfaces often exert a controlling influence on the properties of materials.
Consequently, there is a great desire to engineer these interfaces in order
to optimize beneficial properties and eliminate those that are deleterious.
This optimization is usually achieved by fabrication or processing methods to
control the composition, structure, or crystallography of the interface. This
four day symposium will address these topics in metals, ceramics, and
composites. The discussion will primarily center on internal interfaces,
including grain boundaries, interphase interfaces, and film/coating-substrate
interfaces. The papers will include those which discuss problems which have
been addressed by interfacial engineering and those that discuss the
development of methods of interfacial engineering.

The topics to be covered, as they apply to interfaces, include the following:
o Strength and toughness
o Cohesion and adhesion
o Corrosion and embrittlement
o Fracture and fatigue
o Electrical and optical properties
o Creep and diffusion
o Reactions and mobility
o Measurement and characterization

Partial list of invited speakers : Yet-Ming Chiang (MIT); Gino Palumbo
(Ontario Hydro-electric); Fennell Evans (University of Minnesota); Michael
F. Henry (GE CRD): Vinayak Dravid (Northwestern); Ali Argon (MIT); Kathi
Alexander (ORNL)

Abstracts must be received at MRS Headquarters no later than June 21, 1996 and
must follow the standard MRS abstract model. Abstract templates and
additional information on the meeting can be obtained from MRS (E-mail:
fall96-at-abstracts.mrs.org; WWW: http://www.mrs.org; Phone: 412-367-3003;
Fax: 412-367-4373) or from the symposium organizers.

Symposium Organizers:

Clyde L. Briant
Div. of Engineering
Brown University
PO Box D
Providence, RI 02912
Phone: (401) 863-2626
Fax: (401) 863-7677
E-mail: briant-at-engin.brown.edu

C. Barry Carter
Dept. of Chem. Eng. & Mat. Sci.
Univ. of Minnesota
421 Washington Ave., SE
Minneapolis, MN 44544
Phone: (612) 625-8805
Fax: (612) 626-7246
E-mail: carter-at-cems.umn.edu

Ernest L. Hall
GE CRD
PO Box 8
Room K1-2C27
Schenectady, NY 12301
Phone: (518) 387-6677
Fax: (518) 387-6972
E-mail: hallel-at-crd.ge.com






From: paulc-at-gps.caltech.edu (Paul K. Carpenter)
Date: Fri, 19 Apr 1996 11:05:01 -0700
Subject: Re: Detection of Boron

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: paulc-at-mail.gps.caltech.edu
Message-Id: {v02110100ad9d7c3ca7b9-at-[131.215.67.110]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

On the subject of Boron analysis:

} Mark Darus wrote:
}
} My EDX is a TN-5500 and I'm trying to determine if it is sensitive,
} enough to detect Boron. I have some standards, and in a moment I'm
} going to set up some boric acid and look into it. Just placing it in
} the chamber and getting the counts up, plus switching it to thin window
} doesn't give me overwhelming success from an SPI metal standard that I
} have. I'm sure there is much more to it than that. Is there any advice
} that you may have to offer. My beam current settings are 1, 5, 10, 20
} & 30 KeV.

Just a few comments about the composition of the detector window material
on EDS detectors. The mass absorption coefficients (mac) for Boron Ka
radiation by different absorbing elements are (these values from table 14.3
of Goldstein et al):

Absorber mac
H 1,723
Be 69,937
B 2,861(?)
C 5,945
N 10,118
O 15,774

So B Ka is strongly absorbed by a Be window on your EDS detector, and this
(as you know) is the reason that B Ka cannot be detected with this normal
window material. Note that carbon is fairly transparent to B Ka, either as
an EDS window material via a diamond window, or as a conductive coating
layer on your (nominally non-conductive) specimen. You should be able to
detect B Ka on a carbon-coated sample and/or with a diamond window just
fine; too thick a carbon coat will absorb just about anything, but
comparatively it is not too bad. If you have a BN window, then the
absorption is a little stronger than for the diamond window (average the
macs for B and N). Even slightly worse is for the case of an ice layer
(think of it as a layer of oxygen).

Now if you have a too thick carbon coat, and a thick layer of ice on the
detector crystal, and maybe you are using a thin Be window or a thicker BN
window, then B Ka detection is going to be difficult. As mentioned before,
reconditioning the EDS detector to get rid of the ice layer is mandatory
before attempting to do light element work like this.

Secondly, the pulse processor has to be optimized for processing low energy
x-ray pulses -- if the discriminator is set too high you are filtering out
all the low energy x-ray pulses, and if the time constant is set
incorrectly you may not be getting the best throughput on the detector.
Refer to the manuals at this point.

Suffice to say that optimization of the detector should be done on a pure
element (lump boron) before trying to look at lower levels in your samples.

It was also pointed out by Chuck Garber:
} There is just no better "standard" than a solid lump of pure boron! It
} is inert and it won't be changed by the electron beam. Boric acid will
} be unstable and as others pointed out, it will sublime causing all
} kinds of other problems.

Note however, from the mac values above, that B Ka is strongly absorbed by
many materials (for example, the mac for B Ka by Silicon is 83,702), so you
may see very strong absorption in a B-Si alloy (in fact, you may not
observe a peak at all for lower concentrations of boron because it is
absorbed entirely within the sample!). In this example, the absorption
correction via the "A" part of the ZAF correction is very high for your
sample, compared to your standard (assuming you used pure boron). The
bottom line here is that you absolutely *must* use a standard that is as
close as possible to your samples for good quantitative correction. Of
course, the available boron standards are pure boron, boron nitride (which
is typically polycrystalline and is not a good conductor; single crystal BN
is much better but hard to find in pieces larger than a few microns), and
numerous borosilicate glasses (which of course have the problem of
absorption by silicon). For EDS work you probably have to use pure boron
to avoid peak overlaps, but for WDS analysis things like BN can be used.

The magnitude of the absorption correction is directly related to the
accelerating voltage used, so if you can do analysis at 10 KV rather than
15 KV, you are better off. You can work at lower voltages (like your 5 KV
setting), but the sample surface must be really clean, and you then get
into problems where aspects of the correction algorithms are not really
valid for this range of accelerating voltages.

As a final note, it is relatively meaningless to look at EDS spectra of
pure elements when shopping for an EDS system to do light element analysis.
It is the performance on typical multi-element samples that really
matters. For example, I was looking at the possibility of studying
synthetic vs. natural emeralds (Be3Al2Si6O18) including inspection of the
Be Ka peak by EDS. Even with a freshly reconditioned detector on *open
window* mode, with the SEM probe current full up, no Be peak was observed.
So Be measurement was just not meant to be on an EDS-SEM system. It turns
out that even by WDS, using a Mo4C analyzing crystal (red-hot for Be and
B), and the probe current full up on our microprobe (400 nA -at- 10KV), and
analyzing at low accelerating voltage, and a clean high vacuum turbo-pumped
system, that Be in emerald is barely detectable (count times of up to an
*hour* were used!). I'll just say that boron analysis is a walk in the
park compared to beryllium analysis...

Anyway, good luck!

Paul Carpenter


+----------------------------------------------------+
| Paul K. Carpenter paulc-at-gps.caltech.edu |
| Division Analytical Facility |
| Geological and Planetary Sciences MC 170-25 |
| California Institute of Technology |
| Pasadena, CA 91125 |
| 818-395-6126 (X-ray Lab) 818-568-0935 (Dept. FAX) |
+----------------------------------------------------+






From: K1JIA-at-VAXC.STEVENS-TECH.EDU
Date: Wed, 17 Apr 1996 20:03:39 -0500 (EST)
Subject: UNSUBSCRIBE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Return-path: {Microscopy-request-at-Sparc5.Microscopy.Com}
Received: from Sparc5.Microscopy.Com by VAXC.STEVENS-TECH.EDU
(PMDF V4.3-10 #2500) id {01I3NXCFXVCG9I4ROQ-at-VAXC.STEVENS-TECH.EDU} ; Wed,
17 Apr 1996 22:27:05 -0500 (EST)
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11)
id TAA11158 for dist-Microscopy; Wed, 17 Apr 1996 19:02:01 -0500
Received: from VAXA.STEVENS-TECH.EDU (vaxa.stevens-tech.edu [155.246.1.2])
by Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id TAA11155 for
{microscopy-at-Sparc5.Microscopy.Com} ; Wed, 17 Apr 1996 19:01:59 -0500
Received: from VAXC.STEVENS-TECH.EDU by VAXC.STEVENS-TECH.EDU
(PMDF V4.3-10 #2500) id {01I3NSC2BR749I4R4Y-at-VAXC.STEVENS-TECH.EDU} ; Wed,
17 Apr 1996 20:03:39 -0500 (EST)

Unsubscribe




From: MikGu-at-mme.liu.se (Mikael Gustafsson)
Date: Sun, 21 Apr 1996 11:52:07 +0200
Subject: Hose

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I would like to know if any one has investigated gas permeability properties
of plastic/rubber hose. I'm trying to set up a microscope perfusion system
with flow from a solution with CO2 at a specific concentration. The problem
is that the silicon rubber hose used for interconnection between the parts,
cannot keep CO2, which very quickly escapes into surrounding air.

Has anyone heard of a gas impermeable hose (inner diameter ca 1 mm or less) or


Best regards


=============================================
Mikael Gustafsson MD, PhD
Dept Med. Microbiology and
Dept Internal Medicine, Cardiology section
University Hospital of Linkoping
S 581 85 LINKOPING
SWEDEN

E-Mail: MikGu-at-mme.liu.se
FAX: 046/13/224789
Phone: 046/13/224783
=============================================





From: Brett W. Cockman :      bcockman-at-uniwa.uwa.edu.au
Date: Mon, 22 Apr 1996 09:29:16 +0700 (WAST)
Subject: RE: Semi-thin sections.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Randi,
I suggest transferring your semithin sections to a water bath at about 50
Cand pick up on acid clean slides (this is most important!). Dry on a hot
plate and then place in hot oven (about 60 C +) for at least an hour. This
will guarantee that your sections stay on the slide. I further suggest you
remove the resin by treatment with potassium ethoxide (saturated KOH in
ethanol; make up at least 24 hours before use then filter):
1. Treat with K-ethoxide for about 3 - 4 mins.
2. Wash well with 3 changes of absolute ethanol.
3. Hydrate through 95%, 70% to distilled water and then dry.
4. Stain with toluidine blue or any other stain!

If you need more details please ask.

.
.
Brett W. Cockman
Technologist in Charge
School of Dentistry
University of Western Australia
Voice: (619-2205834)
Fax: (619-2213829)
e-mail; bcockman-at-uniwa.uwa.edu.au




From: Donna Wagahoff :      DWAGAHOF-at-wpsmtp.siumed.edu
Date: Mon, 22 Apr 1996 10:30:33 -0600
Subject: venting print processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We just purchased a new Mohr-Pro print processor but have not been
able to use it because our building engineer is concerned that we must
vent the unit. The manual does not address this issue and there is no
outlet or attachment for venting the chemistry fumes. We have had
several processors over the years, with no exhaust vent and have had
no problems. Would you please share with us whether you have your
print processors connected to a vent system, how you accomplished it,
if it is an OSHA requirement, if your system came with an outlet made
specifically for venting chemistry fumes and any other pertinent
information. Thanks for your assistance.

Donna Wagahoff
SIU School of Medicine
Springfield, Il.
217-782-0898
fax 217-524-3227





From: moxtek-at-MOXTEK.WIN.NET (Clark Turner)
Date: Mon, 22 Apr 1996 09:13:08
Subject: Re: Boron analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Mailer: WinNET Mail, v2.51
Message-ID: {67-at-moxtek.win.net}
Reply-To: moxtek-at-MOXTEK.WIN.NET (Clark Turner)
To: darus-at-cle.dnet.ge.com, microscopy-at-Sparc5.Microscopy.Com

Mark,

A good sample to test the detector capability is solid boron
nitride. It not only gives you the boron K-alpha peak, but also
the nitrogen K-alpha for reference.

D. Clark Turner
Director, Thin Film Products Group
MOXTEK, Inc.
452 West 1260 North
Orem, Utah 84057
phone (801) 225-0930
email moxtek-at-moxtek.win.net



}
}
}
} Hello,
} My EDX is a TN-5500 and I'm trying to determine if it is sensitive,
} enough to detect Boron. I have some standards, and in a moment I'm going to
} set up some boric acid and look into it. Just placing it in the chamber and
} getting the counts up, plus switching it to thin window doesn't give me
} overwhelming success from an SPI metal standard that I have.
} I'm sure there is much more to it than that. Is there any advice that you may
} have to offer. My beam current settings are 1, 5, 10, 20 & 30 KeV.
}
} Mark Darus
}
} Darus-at-cle.dnet.ge.com
}





From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Mon, 22 Apr 1996 09:51:49 -0700 (PDT)
Subject: Re: Semi-thin sections.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Randi,

As already suggested, cleaning the slides is important. For different
plastics, soaking slides for 30 min. in 50% bleach (commercial bleach
mixed 1:1 with dstilled water), or in 70% ethanol containing 1% HCl, will
do the trick instead of messing about with chromic acid. Wash several
times with destilled water after cleaning.

What are you coverslipping with? Some mountants cause wrinkling unless
slides are very clean or the mountant is cut 1:1 with xylene.

With the above approaches, we get reliably flat sections and etch only
when necessary for immunocytochemistry.



Glen MacDonald
Research Scientist
Hearing Research Laboratories of the
Virginia Merrill Bloedel Hearing Research Center
Box 35-7923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156
glenmac-at-u.washington.edu


On Sat, 20 Apr 1996, Randi Olsen wrote:

} Hello,
} I wonder if there is any help somewhere out there:
} Lately we have been fighting some catfish larvaes, embedded
} in Epon/Araldite. Because we need sections of the whole
} length og larvaes up to 4 weeks old, we need big semithin
} sections, and they dont want to strech properly. I have
} small foldes (or waves) in the sections wisible at x40 that
} make it impossible to get sharp photographs at this
} magnification. (The foldes also trap some air between the
} glass slide and the sectiokn). Is there anything we can do
} to make the sections flatter?
}
} Thanks in advance.
}
} Randi Olsen
} Department of Electron Microscopy
} University of Troms=F6
} Norway
}
}
}





From: South Bay Technology :      73531.1344-at-CompuServe.COM
Date: 22 Apr 96 14:55:49 EDT
Subject: Backside Emission Microscopes?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am trying to collect information from the various manufacturers of Backside
Emission Microscopes. If anyone can provide me with names and phone numbers of
appropriate manufacturers, I would greatly appreciate it.

Best regards-

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com
sbt-at-msa.microscopy.com

Manufacturers of Precision Sample Preparation Equipment and Supplies for
Metallography, Crystallography and Electron Microscopy.





From: South Bay Technology :      73531.1344-at-CompuServe.COM
Date: 22 Apr 96 14:57:29 EDT
Subject: ASTM Home Page?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can anyone provide me with the URL for the ASTM home page?

Thank you!

Best regards-

David Henriks TEL: 800-728-2233 (toll-free in USA)
South Bay Technology, Inc. 714-492-2600
1120 Via Callejon FAX: 714-492-1499
San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com
sbt-at-msa.microscopy.com

Manufacturers of Precision Sample Preparation Equipment and Supplies for
Metallography, Crystallography and Electron Microscopy.





From: EmLab :      EMLAB-at-OPUS.MCO.EDU
Date: Mon, 22 Apr 1996 16:00:41 -0500 (EST)
Subject: Re: venting print processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Donna,

Why does the engineer think the processor needs to be vented? There is very
very little mixture of chemicals (ie ammonia fumes). If he/she is so concerned
let them design and make a vent for it. The least that should be done is to
make sure there is enough room ventalation. I myself you not worry.

Best of Luck,
Ed Calomeni
Medical College of Ohio
Toledo, OH 43699
emlab-at-opus.mco.edu




From: EmLab
Date: 4/22/96 2:31 PM
Subject: Re: venting print processor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

"microscopy-at-Sparc5.Microscopy.Co" {microscopy-at-Sparc5.Microscopy.Com}
X-Mailer: Mail*Link SMTP/QM 3.0.0

Reply to: RE} } venting print processors

Donna,

Our safety people required us to have an emergency exhaust fan (with a manual
switch) installed in the darkroom in which the processor is located. Other
than occasional testing, the fan has not been operated in the ten years since
it was installed.

Mike O'Keefe
National Center for Electron Microscopy
Berkeley, CA 94720
--------------------------------------

Why does the engineer think the processor needs to be vented? There is very
very little mixture of chemicals (ie ammonia fumes). If he/she is so
concerned let them design and make a vent for it. The least that should be
done is to make sure there is enough room ventalation. I myself you not
worry.

Best of Luck,
Ed Calomeni
Medical College of Ohio
Toledo, OH 43699
emlab-at-opus.mco.edu






From: slc6-at-lehigh.edu (Sharon Coe)
Date: Tue, 23 Apr 1996 15:40:41 -0400
Subject: Biological SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: slc6-at-mail.Lehigh.EDU
Message-Id: {v01510101ada2e02f1879-at-[128.180.55.59]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Please post the following on the Microscopy Server:

Biological SEM and X-ray Microanalysis

The 1996 Lehigh Short Course, in addition to teaching the theory and basic
prractical skills associated with scanning electron microscopy and x-ray
microanalysis, also provides a thorough understanding of how specimens
should be prepared for imaging and analysis.

Special emphasis is placed on providing answers to the problems associated
with the preparation, microscopy, and analysis of biological, organic,
hydrated and bio-medical materials such as plastics, paints, polymers,
elastomers, insulators, cements, resins, lubricants, pharmaceutical agents,
foods and dairy products, textiles, paper, archeological remains and the
whole spectrum of plant, animal and human tissues. The very nature of
these samples, which are generally beam-sensitive, frequently wet, and
invariably hetereogeneous, require an appreciation and understanding of a
wide range of new techniques and methods. The Lehigh Short Course is
uniquely placed to provide the necessary information and hands-on
instrumentation to address these problems.

Specific instruction will be given in low temperature and low voltage
microscopy and analysis; sample preparation and specimen coating; immuno-
and histo-chemical and staining techniques to localize specific chemical
ligands; non-invasive preparative methods and techniques for diffusible and
soluble substances; environmental microscopy and analysis, and the
strategies and tactics to employ when examining and analysing
beam-sensitive samples. Special biological lectures by Dr. Patrick Echlin,
University of Cambridge.

For a free brochure contact:

Sharon Coe
Materials Science Department
Lehigh University
5 East Packer Avenue
Bethlehem, PA 18015
Phone: 610/758-5133
Fax: 610/758-4244
e-mail: slc6-at-lehigh.edu






From: IAN HALLETT :      ihallett-at-MARCCRI.MARC.CRI.NZ
Date: Tue, 23 Apr 1996 09:00:32 GMT+1200
Subject: Re: venting print processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {m0uBVpX-0003L7C-at-fast.net}

Dear Donna

We have an externally venting overbench extraction hood over both
the print processor and an adjacent wet area. This appears to work
well and certainly eliminates the smell of the warm chemicals.

All photographic processing should be carried out with some form of
fume extraction, something that is often ignored. Our photographic
section installed such hoods in their darkrooms 12 or 13 years ago
after one of the photographers developed respiratory problems
carrying out conventional hand printing..

Ian Hallett

(Donna Wagahoff writes ...)
} We just purchased a new Mohr-Pro print processor but have not been
} able to use it because our building engineer is concerned that we must
} vent the unit. The manual does not address this issue and there is no
} outlet or attachment for venting the chemistry fumes. We have had
} several processors over the years, with no exhaust vent and have had
} no problems. Would you please share with us whether you have your
} print processors connected to a vent system, how you accomplished it,
} if it is an OSHA requirement, if your system came with an outlet made
} specifically for venting chemistry fumes and any other pertinent
} information. Thanks for your assistance.
}
} Donna Wagahoff
} SIU School of Medicine
} Springfield, Il.
} 217-782-0898
} fax 217-524-3227

**************************************************


Ian Hallett
HortResearch
Mt Albert Research Centre
Private Bag 92 169
Auckland, New Zealand
Fax 64-9-815 4201
Telephone 64-9-849 3660
EMail ihallett-at-hort.cri.nz




From: Brett W. Cockman :      bcockman-at-uniwa.uwa.edu.au
Date: Tue, 23 Apr 1996 10:42:21 +0700 (WAST)
Subject: RE: Semi-thin sections.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-NUPop-Charset: English

Hello Robert,
In answer to your questions:
1. I usually use a cotton bud to pick up the sections and 'roll' them onto
the surface of the waterbath.
2. I havn't used spurr's resin or mixtures with other epoxy's but it is
worth a try!.
3. I add KOH ot cover the base of a 1lt reagent bottle (screw top) and fill
with ethanol (absolute). The solution goes brown after a day or so , then
filter before use. It keeps reasonably well. Please note that it is very
corrosive!!!.
4. The original reference is: Imai, Y et al (1968): Removing method of resin
for histological section following halogenation. Electron Microsc. vol 17,84

Regards,

Brett.




From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Mon, 22 Apr 1996 17:16:20 -0600
Subject: Central States Microscopy Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130501ada1be9cfcbc-at-[131.230.97.69]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

******************************************************************************

ENJOY MICROSCOPY AND THE GREAT OUTDOORS - WHAT A COMBINATION!

******************************************************************************

The Spring meeting of the Central States Microscopy Society will take place
on June 21, 1996 in Carbondale, IL at the Giant City State Park Lodge. The
theme of the meeting will be "Ride the Wave" (e.g., technological wave).
Innovative uses of new and established technologies (microwave, LM, SEM,
TEM, scanned probes, etc) are invited. Biological and physical science
presentations are welcome.

Cash prizes for best student presentations. Typically this ranges from
$50-100, depending upon number of entrants and quality of work presented.
Abstracts must be received 3 weeks prior to meeting to be considered in the
student competition. Follow MSA abstract guidelines.

CSMS corporate sponsors are invited to participate by means of talks and
demonstrations of equipment. Contact me well in advance of the meeting,
however, if you require tables, space or special electrical setups. There
will be no charge for CSMS corporate sponsors.

Programs, lodging recommendations, directions to meeting site will be sent
several weeks in advance of the meeting.

Speakers and presentors: please contact John Bozzola for topics/timing no
later than May 21, 1996.


******************************************************************************






#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Tue, 23 Apr 1996 17:57:23 +1100
Subject: Re: venting print processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: st004718-at-brandywine.otago.ac.nz
Message-Id: {v01540b0aada22b81fb0d-at-[139.80.120.139]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We have a Rapiline print processor which came with a (very small) built-in
fan to extract fumes from the tanks... and blow them into the room. The
installation did not include connecting this to any extract system although
the technician indicated that this could be done. Unlike your experience
however, we found the fumes unpleasant so we had our workshop make elbows
and ducts to connect the output to the building extract sustem and we have
noticed that the air is much fresher in the darkroom as a result. I don't
know if the fumes were unsafe before but it is certainly more pleasant to
use be in that room now and we feel that it was a worthwhile modification.
I suppose whether or not is worthwhile depends a lot what photographic
chemistry you are using and at what temperature (we use 27degC which is
relatively cool). I wouldn't assume that because the manual doesn't mention
it that it isn't worth doing, my expectation of manuals gets lower with
every new bit of equipment we buy.

Best regards,

Richard


Richard Easingwood
South Campus Electron Microscope Unit
School of Medical Sciences
University of Otago
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

SOUTHERNMOST E.M UNIT IN THE WORLD















From: Bo Johansen :      BOJ-at-bot.ku.dk
Date: Tue, 23 Apr 1996 08:11:24 GMT+0200
Subject: Mounting medium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello fellow microscopists

I need a little help.
I am using the NBT/BCIP method to detect alkaline phosphatase in
indirect immunomarking on plant tissue sections. I want to
mount these sections permanently and I have tried DPX but the reduced
tetrazolium salt just (re)crystallize and any specific staining is
lost. Any suggestion about what to use instead of DPX is much
appreciated.

Thank you

Bo

_____________________________________________________________________
Bo Johansen E-Mail: BoJ-at-bot.ku.dk
Botanical Institute Vioce: +45 3532 2157
Gothersgade 140 FAX: +45 3313 9104
DK-1123 Copenhagen K, Denmark http://www.bot.ku.dk/www/staff/boj.htm
---------------------------------------------------------------------






From: nano-at-ns1.LIHTI.ORG (Nanoprobes Inc.)
Date: Tue, 23 Apr 1996 08:43:50 -0500
Subject: Independent Sales Representatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604231241.IAA25293-at-lihti.org}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Nanoprobes, Inc is looking for independent sales representatives to carry their
immunoreagent product line. This position(s) are commission only. Please send
inquires/resumes to Ms. Tempel at Nanoprobes, Inc. 25 E. Loop Rd. Ste 124,
Stony Brook NY 11790-3350,Phone: 516 444 8815 or Fax 516 444 8816;
E.mail: nano-at-mail.lihti.org.








From: Mev H van der Merwe :      HVDM-at-op1.up.ac.za
Date: Tue, 23 Apr 1996 15:04:29 GMT+2
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


please unsubscribe




From: nano-at-ns1.LIHTI.ORG (Nanoprobes Inc.)
Date: Tue, 23 Apr 1996 10:42:13 -0500
Subject: Independent Sales Representatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604231440.KAA25783-at-lihti.org}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Nanoprobes, Inc is looking for independent sales representatives to carry their
immunoreagent product line. This position(s) are commission only. Please send
inquires/resumes to Ms. Tempel at Nanoprobes, Inc. 25 E. Loop Rd. Ste 124,
Stony Brook NY 11790-3350,Phone: 516 444 8815 or Fax 516 444 8816;
E.mail: nano-at-mail.lihti.org.











From: olinsal-at-bio.ornl.gov (Ada L. Olins)
Date: Tue, 23 Apr 1996 10:41:49 +0500
Subject: RE: Semi-thin sections.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v02110101ada21882160c-at-[134.167.112.26]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I have good success picking up semi-thin sections with a fine wire loop
that I attached to a wooden stick. With a high meniscus approach the
section from below and lift up. The section can then be deposited on a
water droplet (on a slide) either by inverting the loop on the water
droplet or by floating the section off. Heat slide GENTLY to avoid
wrinkles and dry the section.

For em: I usually transfer the section to a beam capsul filled with
water, and put it in a warm oven (at 37-45 deg.C), to remove the wrinkles.
I then use a plastic film-(e.g. parloidin) coated grid to pick up the
section. This prevents "electrostatic jumping" of the section when it is
dry.

Good luck, Ada

ADA L. OLINS
THE UNIVERSITY OF TENNESSEE-OAK RIDGE and
THE BIOLOGY DIVISION, ORNL
P.O. BOX 2009 TEL: 423 574 1269
OAK RIDGE, TN 37831-8077 FAX: 423 574 1274








From: ENERGY BEAM SCIENCES, INC :      75767.640-at-CompuServe.COM
Date: 23 Apr 96 10:35:37 EDT
Subject: Re: manuals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Richard Easingwood wrote something in his recent post regarding ventilation and
film processors that struck a raw nerve. He wrote:
"I wouldn't assume that because the manual doesn't mention it that it isn't
worth doing, my expectation of manuals gets lower with every new bit of
equipment we buy."

We have just received CE (European Community) certification for our laboratory
microwave processor after many months and many thousands of dollars. CE
compliance with electrical safety regulations has been required since Jan 1,
1996, for *all* electrical equipment exported into Europe.

Many of the CE regulations relate to what must be in the users' manual.
Everything that is required for the manual must be translated into several
languages, in addition to conforming to complicated rules regarding content and
format. The end result is that we've cut our very thorough manual of almost 200
pages down to fewer than 20 pages at the urging of our CE certification
consultant.

I would expect that the trend that Richard points out in his post will continue
to worsen as more international manufacturing companies deal with the new CE
regulations.

Steven Slap, Vice-President
ebs-at-ebsciences.com





From: BARBARA.HARTMAN-at-smtpgw.inet.spri.sp.com (BARBARA HARTMAN)
Date: Tue, 23 Apr 1996 14:36:02 -0400
Subject: General Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Protocols in:

Microscopic Imaging, Immunocytochemistry and Image Analysis

Twenty-Second Annual Program

The George Washington University Center for Microscopy
and Image Analysis

June 4-7, 1996

For further course information, please contact:

Mr. Fred Lightfoot
202-994-2881 (PHONE)
202-994-8885 (FAX)
E-Mail: FredL-at-INDY.CMIA.GWUMC.EDU




From: MELLIOTT-at-prl.pulmonary.ubc.ca
Date: Tue, 23 Apr 1996 12:12:22 +0800PST
Subject: anti-myeloperoxidase

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a source for antibodies to rabbit myeloperoxidase??
Can't find them in Linscott's or the MSRS catalogue. Do any of the
antibodies to human cross-react with rabbit??

Thanks
Mark Elliott
Pulmonary Research Lab, UBC
Vancouver Canada





From: cal-at-ssnet.com (cal Montgomery)
Date: Tue, 23 Apr 1996 16:00:11 -0400
Subject: Lehigh course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Have a customer interested in EM courses. I suggested the Lehigh course.
Would like to know who to have her contact to get info on the Lehigh course.
TIA
Cal Montgomery





From: Carmine M. Pariante :      cparian-at-emory.edu
Date: Tue, 23 Apr 1996 17:30:13 -0400 (EDT)
Subject: Lehigh course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


please unsuscribe





From: Gary H. Zajic :      zajic-at-umich.edu
Date: Tue, 23 Apr 1996 15:16:17 -0400 (EDT)
Subject: blocking endogenous peroxidase

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello, I am doing immunocytochemical experiments on mouse cochlea
using frozen sections. I find that I must block endogenous peroxidase
in these tissues. Most protocols I see require dehydration to 100%
EtOH or MeOH when blocking with H2O2. Cochlear tissue is very delicate
and there is a risk losing some of the fine structure with these
additional steps. My questions are: why can't I just block with H2O2
in my buffer? is the alcohol necessary for penetration allowing better
access of the H2O2 to the tissue? These are 10 micron sections.
Thanks in advance. Sincerely, Gary


zajic-at-umich.edu




From: Bob Citron :      Bob_Citron-at-cc.chiron.com
Date: 4/22/96 1:56 PM
Subject: TEM Contract Lab

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Anyone out there know of a lab where we can get some TEM work done near the
Irvine, California
area. Within Orange County will do. Thanks to all.





From: Louisa.Howard-at-Dartmouth.EDU (Louisa Howard)
Date: 23 Apr 96 13:19:05 EDT
Subject: Long term fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {13739592-at-prancer.Dartmouth.EDU}

Our lab has been asked to help on a possible project studying mammalian brain
tissue. The material will be collected in the field( a few thousand miles fom
our lab) and storage times could be up to 3-4 weeks, before the material
reaches the lab for processing. I have an excellent fixation method for brain
tissue, but its fairly involved and the solutions must be made up fresh; not a
real possibility for this field work. In the past, I have had great success,
using PIPES buffer for the glutaraldehyde fixation, whenever mammalian em
material had to be stored for 3-4 days at 4 degrees C. I would make up the
fixative bottle of 4% GTA in 0.15M PIPES, pH 7.35 and send it out. They would
add their samples and mail(at -at-4 degrees C) them to me. I've also had suceess
using the GTA/PIPES fixation and then transferring samples to a buffer of 0.1M
Na Cacodylate with 0.15M Sucrose, pH 7.4. and holding the sample for a week,
at 4 degrees C.
Has anyone had experience in keeping a sample for 4 weeks before final
processing. Would it be better to store the sample:
1. in PIPES buffer, after GTA/PIPES fixation?
2. 0.1M Na Cacodylate with 0.15M Sucrose, pH 7.4. after GTA/PIPES fixation?
3. leave in GTA/PIPES fixative for the entire time at 4 Degrees C. ( I worry
about the staibility of the GTA over such a long time)
4. Some other method, that works with mammalian tissue.

Any suggestions or thoughts on long term fixation/storage would be greatly
appreciated.
thanks
Louisa Howard




From: becks-at-sunynassau.edu (Steve Beck)
Date: Tue, 23 Apr 1996 19:20:34 -0500
Subject: Summer TEM Course Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


SUMMER COURSE ANNOUNCEMENT - Transmission Electron Microscopy (BIO. 221)

NASSAU COMMUNITY COLLEGE

A five week course in Biological Transmission Electron Microscopy is being
offered by the Biology Department of Nassau Community College. This is a 4
credit course offered over the first summer session, between May 28 and
June 27, 1996. The class meets from 8:00 am to Noon four days per week
(usually Monday thru Thursday).

This is a "hands-on" course emphasizing biological specimen preparation,
ultra-thin sectioning involving block trimming, glass knifemaking and
operation of the ultramicrotomes (Sorvall MT-2B and LKB Ultrotome III),
thick and ultra-thin section mounting and contrast staining (UA and Pb
citrate), grid support films (formvar, carbon), student operation of the
TEM (Hitachi HS-8, Philips EM 300) and production of electron micrographs
through the process of black & white photography, and electron micrograph
analysis. Students will work on a chosen sample(s) with the goal of
producing a portfolio of high quality TEM photomicrographs of that
sample(s).

The course is widely transferrable and the cost per credit is reasonable at
$78 per credit.

Registration for summer courses can be conducted via phone by calling (516)
572-7131 or (516) 572-7372 or (516) 572-7425, Monday through Thursday
2:30pm to 7:00pm (This phone service ends on May 2, 1996)

More information about summer registration and course offerings is
available at our web site {http://www.sunynassau.edu}

The catalog description is specified below. If you have further questions,
you should e-mail me directly at the address below.

CATALOG DESCRIPTION
BIO 221: Transmission Electron Microscopy 4 cr.
Prerequisites: BIO 109-110 or equivalent, CHE 151-152 or equivalent.
An introduction to the basic principles of transmission electron
microscopy including tissue preparation, microscope (TEM) operation, black
& white photography, and micrograph interpretation. The entire laboratory
is devoted to the development of skills and preparative techniques involved
with the operation of an actual transmission electron microscope.
(3 lecture, 3 laboratory hours). Laboratory fee applies.



Stephen Beck
Bio-Imaging Center/Electron Microscopy
Department of Biology
Nassau Community College
Garden City, NY 11530
Voice Mail: (516) 572-7829





From: knoff-at-lipovx.lbl.gov (Laura Knoff)
Date: Tue, 23 Apr 1996 12:37:01 -0800
Subject: No. Cal Micro. Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604231932.OAA21729-at-Sparc5.Microscopy.Com}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Northern CA Society for Microscopy (NCSM) Meeting Announcement
}
} Where: Roche Biochemicals (formerly Syntex), Gallery
} Conference Center Bldg A-2
} 3401 Hillview Ave, Palo Alto CA
}
} When: Thursday, May 9, 1994, from 2:30 to 8:45 PM
}
}
} 2:30 -3:00 Registration
} Speakers
} 3:00-3:45 Hank W. Bass, UC Berkeley, Molecular and Cellular
} Biology, "Analysis of Meiosis using 3-D
} Deconvolution Light Microscopy". 4:00-4:45 David Blake,
} NASA Ames Research Center, Space Sciences Division
} "Electron Microscopy of Astrophysical Ice"
} 5:00 - 6:30 Social Hour and Local Society A filiate meeting
} sponsored by JEOL Instruments
} 6:30-7:30 Dinner. Choices: Marinated Stripped Loin with sundried
} tomatoes, Lemon chicken with caper sauce, or roasted
} vegetable lasagna,
} each with garlic mashed potatoes, summer squash medley,
} mixed green salad, dinner rolls, and dessert.
} 7:45 - 8:45 Paul Carpenter, Manager Analytical and Geology, Cal
} Institute of Technology, "XRay and Optical
} Methods as Applied to the Study of Gemstones."
}
} Please help us to plan accurately. Make your reservation before Friday MAY
} 3rd
} Fax reservation(s) with meal preference to Laura Knoff at (510) 486-4750.
} E-Mail address = KNOFF-at-LIPOVX.LBL.GOV
} Phone = (510) 486-4088, if leaving a message please speak slowly and clearly.
} Price: $17 for Loin, $14 for Chicken, $13 for Lasagna,$8 for student
} members.
} Avoid standing in line by pre-paying. Send your check, payable to NCSM,
} to NCSM c/o Laura Knoff
} LawrenceBerkeley National Lab
} Bldg 1, Room 264
} 1 Cyclotron Rd
} Berkeley, CA 94720
Sorry about the format (or lack of it).






From: knoff-at-lipovx.lbl.gov (Laura Knoff)
Date: Tue, 23 Apr 1996 12:41:58 -0800
Subject: No. Cal Micro. Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604231937.OAA21742-at-Sparc5.Microscopy.Com}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Date: Tue, 23 Apr 1996 12:37:02 -0800
} To: microscopy-at-Sparc5.Microscopy.Com
} From: knoff-at-lipovx.lbl.gov (Laura Knoff)
} Subject: No. Cal Micro. Meeting
} Cc:
} Bcc:
} X-Attachments:
}
} } Northern CA Society for Microscopy (NCSM) Meeting Announcement
} }
} } Where: Roche Biochemicals (formerly Syntex), Gallery
} } Conference Center Bldg A-2
} } 3401 Hillview Ave, Palo Alto CA
} }
} } When: Thursday, May 9, 1994, from 2:30 to 8:45 PM
} }
} }
} } 2:30 -3:00 Registration
} } Speakers
} } 3:00-3:45 Hank W. Bass, UC Berkeley, Molecular and Cellular
} } Biology, "Analysis of Meiosis using 3-D
} } Deconvolution Light Microscopy". 4:00-4:45 David
} } Blake, NASA Ames Research Center, Space Sciences Division
} } "Electron Microscopy of Astrophysical Ice"
} } 5:00 - 6:30 Social Hour and Local Society A filiate meeting
} } sponsored by JEOL Instruments
} } 6:30-7:30 Dinner. Choices: Marinated Stripped Loin with
} } sundried tomatoes, Lemon chicken with caper sauce, or
} } roasted vegetable lasagna,
} } each with garlic mashed potatoes, summer squash
} } medley, mixed green salad, dinner rolls, and
} } dessert.
} } 7:45 - 8:45 Paul Carpenter, Manager Analytical and Geology, Cal
} } Institute of Technology, "XRay and Optical
} } Methods as Applied to the Study of Gemstones."
} }
} } Please help us to plan accurately. Make your reservation before Friday MAY
} } 3rd
} } Fax reservation(s) with meal preference to Laura Knoff at (510) 486-4750.
} } E-Mail address = KNOFF-at-LIPOVX.LBL.GOV
} } Phone = (510) 486-4088, if leaving a message please speak slowly and clearly.
} } Price: $17 for Loin, $14 for Chicken, $13 for Lasagna,$8 for student
} } members.
} } Avoid standing in line by pre-paying. Send your check, payable to NCSM,
} } to NCSM c/o Laura Knoff
} } LawrenceBerkeley National Lab
} } Bldg 1, Room 264
} } 1 Cyclotron Rd
} } Berkeley, CA 94720
} Sorry about the format (or lack of it).
}






From: knoff-at-lipovx.lbl.gov (Laura Knoff)
Date: Tue, 23 Apr 1996 12:41:58 -0800
Subject: No. Cal Micro. Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604231937.OAA21742-at-Sparc5.Microscopy.Com}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Date: Tue, 23 Apr 1996 12:37:02 -0800
} To: microscopy-at-Sparc5.Microscopy.Com
} From: knoff-at-lipovx.lbl.gov (Laura Knoff)
} Subject: No. Cal Micro. Meeting
} Cc:
} Bcc:
} X-Attachments:
}
} } Northern CA Society for Microscopy (NCSM) Meeting Announcement
} }
} } Where: Roche Biochemicals (formerly Syntex), Gallery
} } Conference Center Bldg A-2
} } 3401 Hillview Ave, Palo Alto CA
} }
} } When: Thursday, May 9, 1994, from 2:30 to 8:45 PM
} }
} }
} } 2:30 -3:00 Registration
} } Speakers
} } 3:00-3:45 Hank W. Bass, UC Berkeley, Molecular and Cellular
} } Biology, "Analysis of Meiosis using 3-D
} } Deconvolution Light Microscopy". 4:00-4:45 David
} } Blake, NASA Ames Research Center, Space Sciences Division
} } "Electron Microscopy of Astrophysical Ice"
} } 5:00 - 6:30 Social Hour and Local Society A filiate meeting
} } sponsored by JEOL Instruments
} } 6:30-7:30 Dinner. Choices: Marinated Stripped Loin with
} } sundried tomatoes, Lemon chicken with caper sauce, or
} } roasted vegetable lasagna,
} } each with garlic mashed potatoes, summer squash
} } medley, mixed green salad, dinner rolls, and
} } dessert.
} } 7:45 - 8:45 Paul Carpenter, Manager Analytical and Geology, Cal
} } Institute of Technology, "XRay and Optical
} } Methods as Applied to the Study of Gemstones."
} }
} } Please help us to plan accurately. Make your reservation before Friday MAY
} } 3rd
} } Fax reservation(s) with meal preference to Laura Knoff at (510) 486-4750.
} } E-Mail address = KNOFF-at-LIPOVX.LBL.GOV
} } Phone = (510) 486-4088, if leaving a message please speak slowly and clearly.
} } Price: $17 for Loin, $14 for Chicken, $13 for Lasagna,$8 for student
} } members.
} } Avoid standing in line by pre-paying. Send your check, payable to NCSM,
} } to NCSM c/o Laura Knoff
} } LawrenceBerkeley National Lab
} } Bldg 1, Room 264
} } 1 Cyclotron Rd
} } Berkeley, CA 94720
} Sorry about the format (or lack of it).
}






From: Greg2NJ-at-aol.com
Date: Tue, 23 Apr 1996 21:02:15 -0400
Subject: Kidney Fixation for Routine TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings:

Am interested in finding out the best overall fixation for Kidney (medulla
and cortex).
Interested in good preservation (who isnt) as well as defined membrane
boundries. Specifically looking at mitochondria.






From: p&s-at-ultra.net.au (Probing & Structure)
Date: Wed, 24 Apr 1996 13:03:13 +1000
Subject: Home Page Probing & Structure

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Microscopists & Microanalyst:

Probing & Structure has a large fully illustrated microscopy
catalogue on the
www. Prices are shown in Australian dollars. Most items are shipped FIS.
Currency converter, numerous good links and many MSDS documents are also
available. It's worth a bookmark

http://www.ultra.net.au/~pns/index.html

Jim Darley





From: dianavd-at-eye.usyd.edu.au (Diana van Driel)
Date: Wed, 24 Apr 1996 14:38:39 +1100
Subject: Re: Long term fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: diana-at-pc-0.eye.usyd.edu.au
Message-Id: {v01540b03ada3511abe4d-at-[129.78.203.31]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

}
} Any suggestions or thoughts on long term fixation/storage would be greatly
} appreciated.
} thanks
} Louisa Howard

I have had no problems with long term (years) storage of retina in
cacodylate buffer (no additives) after glutaraldehyde fixation. Long term
(weeks) storage in glut/cacodylate is also not a problem - the tissue
doesn't harden excessively - though I've never done it with brain. All at 4
deg C of course.

Diana van Driel
Dept Ophthalmology
Sydney University
AUSTRALIA 2006






From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Wed, 24 Apr 1996 07:53:02 +0100 (BST)
Subject: Re: Lehigh course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to Cal Montgomery and any other people interested in EM (SEM)
courses they should contact Sharon Coe tat e-mail address
{slc6-at-Lehigh.EDU} who will be able to give you details of the all
singing, all dancing, bells and whistles LEHIGH SHORT COURSE which will
covers everything you want to know and were afraid to ask about SEM and
x-ray microanalysis of every conceivable type of specimen (as can be seen
it does not, alas, cover typing)

Patrick Echlin
Cambridge UKOn Tue, 23 Apr 1996, cal Montgomery wrote:

} Have a customer interested in EM courses. I suggested the Lehigh course.
} Would like to know who to have her contact to get info on the Lehigh course.
} TIA
} Cal Montgomery
}
}




From: Tony Bruton :      bruton-at-emu.unp.ac.za
Date: Wed, 24 Apr 1996 09:56:27 +0200
Subject: Standards for EDX on TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s17dfaf2.020-at-smtp.unp.ac.za}
X-Mailer: Novell GroupWise 4.1

If anyone out there has got advice/recommendations/experience on the
preparation and use of standards for TEM+EDX we would be most
interested to hear from you.

Though our emphasis will ultimately be on biological applications, any
input would be of interest at this stage.

Tony Bruton
Centre for Electron Microscopy
University of Natal, Pietermaritzburg
Private Bag X01, Scottsville 3209
KwaZulu Natal, South Africa

e-mail: bruton-at-emu.unp.ac.za





From: JSmiley-at-nwu.edu (John Smiley)
Date: Wed, 24 Apr 1996 08:42:19 -0600
Subject: Re: blocking endogenous peroxidase

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For postmortem human brain tissue, which is also delicate, we use 1% H2O2
in PBS for 1 hour at 4C. This incubation is done without shaking, because
the peroxide is also somewhat harsh on the tissue.

} }
} } Hello, I am doing immunocytochemical experiments on mouse cochlea
} } using frozen sections. I find that I must block endogenous peroxidase
} } in these tissues. Most protocols I see require dehydration to 100%
} } EtOH or MeOH when blocking with H2O2. Cochlear tissue is very delicate
} } and there is a risk losing some of the fine structure with these
} } additional steps. My questions are: why can't I just block with H2O2
} } in my buffer? is the alcohol necessary for penetration allowing better
} } access of the H2O2 to the tissue? These are 10 micron sections.
} } Thanks in advance. Sincerely, Gary
} }
} }
} } zajic-at-umich.edu






From: Walt Bobrowski (313) 996-7814 :      bobroww-at-aa.wl.com
Date: Wed, 24 Apr 1996 09:32:29 -0400 (EDT)
Subject: Re: Kidney Fixation for Routine TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mr-Received: by mta SRVR05.MUAS; Relayed; Wed, 24 Apr 1996 09:32:29 -0400
Mr-Received: by mta SRVR05; Relayed; Wed, 24 Apr 1996 09:32:30 -0400
Mr-Received: by mta SRVR01; Relayed; Wed, 24 Apr 1996 09:34:22 -0400
Disclose-Recipients: prohibited
MSA MICROSCOPY MAILING LIST {MICROSCOPY-at-Sparc5.Microscopy.Com}
Message-Id: {4129320924041996/A20679/SRVR05/11A4C2601C00*-at-MHS}
X-Envelope-To: Greg2NJ-at-aol.com, MICROSCOPY-at-MSA.MICROSCOPY.COM
Autoforwarded: false
Mime-Version: 1.0
Content-Type: TEXT/PLAIN; CHARSET=US-ASCII
Content-Transfer-Encoding: 7BIT
Importance: normal
Priority: normal
Sensitivity: Company-Confidential
Ua-Content-Id: 11A4C2601C00
X400-Mts-Identifier: [;4129320924041996/A20679/SRVR05]
Hop-Count: 2

Retrograde perfusion fixation via abdominal aorta. See: Griffith, LD et al.
(1967). The ultrastructure of the functioning kidney. Lab. Invest. 16,
220-246.
Best regards,

Walter F. Bobrowski
Subcellular Pathology
Parke-Davis Pharmaceutical Research
Ann Arbor, MI 48105

TEL: 313-996-7814
FAX: 313-996-5001
E-Mail: BOBROWW-at-AA.WL.COM
} Greetings:
}
} Am interested in finding out the best overall fixation for Kidney (medulla
} and cortex).
} Interested in good preservation (who isnt) as well as defined membrane
} boundries. Specifically looking at mitochondria.
}






From: rosemary-at-aec.env.gov.ab.ca (Rosemary K. HARRIS)
Date: Wed, 24 Apr 1996 08:59:50 -0600
Subject: long term storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To expand upon Louisa Howard's question of long term storage -she is
asking about storage in cacodylate buffer.

The question has recently been asked here about long term storage in
phosphate buffers. We have a variety of fish, rat and cow tissues that were
collected (or will be collected) "just in case" EM may be needed. Some of
these tissues have been around for at least 10 years and will likely be here
another 10 years.

What are the current thoughts on storage for long lengths of time?
Would it be better to keep the tissues in fixative as opposed to buffer?
Would cacodylate buffer be preferable to phosphate (or is there something
else)? Any special techniques for processing old tissues and getting useable
results? How long have tissues been kept around and still been successfully
worked with?

Thanks for any advice from all the experts on the list.

Rosemary Harris
Electron Microscopy Laboratory
Alberta Environmental Centre






From: Joiner Cartwright, Jr., Ph.D. :      joiner-at-bcm.tmc.edu
Date: Wed, 24 Apr 1996 09:18:08 -0500
Subject: Re: Kidney Fixation for Routine TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604241519.KAA06313-at-watson.bcm.tmc.edu}
X-Sender: joiner-at-bcm.tmc.edu (Unverified)
X-Mailer: Windows Eudora Version 2.1.1
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

At 09:02 PM 4/23/96 -0400, you wrote:

} Greetings:
}
} Am interested in finding out the best overall fixation for Kidney (medulla
} and cortex).
} Interested in good preservation (who isnt) as well as defined membrane
} boundries. Specifically looking at mitochondria.
}
}
**********************
We get good results fixing renal biopsies in 3% glutaraldehyde in 0.09 M
PIPES, pH 7.2 followed by 2.0% osmium in PIPES. We then enbloc stain with
saturated aqueous UAc for 45 minutes and embed in Spur's resin. The tissue
remains in the glutaraldehyde for varying times up to a day because these
are medical specimens coming from various hospitals.


Joiner Cartwright, Jr., Ph.D.

Director, Electron Microscopy tel.: (713)798-4658
Department of Pathology, Rm.286-A FAX: (713)798-3945
Baylor College of Medicine Internet: joiner-at-bcm.tmc.edu
One Baylor Plaza Compuserve: 71555,1206
Houston, Texas 77030 U.S.A.






From: Joyce Craig :      Chicago.State.University-at-uxa.ecn.bgu.edu
Date: Wed, 24 Apr 1996 11:04:54 -0500
Subject: critical point dryer problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604241708.MAA28581-at-uxa.ecn.bgu.edu}
X-Sender: bafpjec-at-pop3.ecn.bgu.edu
X-Mailer: Windows Eudora Light Version 1.5.2
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We have a Bal-Tec CPD030. It has been used very little. This is because we
cannot get it to drain. It might take all night for the alcohol-CO2 mixture
to empty. The instructions seem very straigt-forward. We fill partially
with 100% ethanol (with or without a sample), cool to 8 degrees C, let in
the carbon dioxide (yes, the tank is the kind with the dip tube), then press
the medium out button. At first it seems to empty, then gets very slow.
Has anyone had a similar problem? The valves seem to work as I hear a loud
click when pressing buttons. The filters are clean. Any other ideas? thanks.





From: MicroWorld News :      spb-at-wwa.com
Date: Wed, 24 Apr 1996 12:30:01 -0500
Subject: Re: Lehigh course and other courses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {2.2.32.19960424173001.0068bc04-at-pop.wwa.com}
X-Sender: spb-at-pop.wwa.com
X-Mailer: Windows Eudora Pro Version 2.2 (32)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

There is a list of microscopy short courses and in-house trainers on the WWW
at http://www.mwrn.com/product/train.htm The list has direct links to
information on the WWW and e-mail addresses. Lehigh information is included.

Susanne Pignolet Brandom
MicroWorld Resources and News
http://www.mwrn.com/
spb-at-wwa.com
847-548-6522

At 04:00 PM 4/23/96 -0400, you wrote:
} Have a customer interested in EM courses. I suggested the Lehigh course.
} Would like to know who to have her contact to get info on the Lehigh course.
} TIA
} Cal Montgomery
}
}
}





From: Larry Ackerman :      mishot-at-itsa.ucsf.edu
Date: Wed, 24 Apr 1996 13:23:11 -0700 (PDT)
Subject: long term storage/fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I recall that some NASA sponsored studies of long term fixation found the
best results leaving tissue in the glutaraldehyde fix. This may even have
been presented at an EMSA meeting in the mid 1970's but I haven't found
the reference yet.





From: William Tivol :      tivol-at-wadsworth.org
Date: Wed, 24 Apr 1996 17:26:52 -0400 (EDT)
Subject: Re: Standards for EDX on TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


}
} If anyone out there has got advice/recommendations/experience on the
} preparation and use of standards for TEM+EDX we would be most
} interested to hear from you.
}
} Though our emphasis will ultimately be on biological applications, any
} input would be of interest at this stage.
}
Dear Tony,
A few years ago, the late, lamented Chuck Fiori produced some
lithium borate glass standards especially designed for average Z to be
equal to that of tissue. I think someone at NIST can tell you more.
Good luck.
Yours,
Bill Tivol




From: Eric Steel :      steel-at-enh.nist.gov
Date: Wed, 24 Apr 1996 12:35:40 -0400 (EDT)
Subject: EDS standards for AEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In reply to Tony Bruton's request for info on EDS/TEM standards:

There is a National Institute of Standards and Technology Standard Reference
Materials (SRM) specifically designed for the X-ray analysis on the TEM,
though not for biological applications, and unluckily it is not free.

Here is a brief description of the material, so that you can see if you
might be interested:

The standard is called "SRM 2063 Microanalysis Thin Film" and consists of a
thin film of sputtered mineral glass supported by a carbon film and a copper
TEM grid. The composition is certified as listed below and was determined
by several independent techniques after the glass was deposited on the
grids. Thus the certified composition values take into account sample
preparation. The thickness of the glass (76 nm) and density (3.1 gm/cm**3)
are reported (though not certified). No preparation is necessary for use in
the TEM.

Table of Concentration Values for SRM 2063 Mineral Glass
Element Concentration
(% wt.)
O 43.2
Mg 7.97
Si 25.34
Ca 11.82
Fe 11.06

The composition allows for a range of relative sensitivity values to be
determined for many commonly analyzed elements and x-ray lines. Thus the
standard can be a primary, traceable way of checking other in-house
standards you may use. The standard is robust under many handling and beam
conditions, though very high beam dose may cause a change in composition (as
noted on the certificate of analysis.) We have used one standard grid for
about seven years in our laboratory to monitor/compare several instruments
and detectors.

For more information about obtaining the standard you may contact:

Standard Reference Materials Program
National Institute of Standards and Technology
Gaithersburg, MD 20899-0001
USA

Phone: 301-975-6776
FAX: 301-948-3730
e-mail: SRMINFO-at-enh.nist.gov





From: Paul Webster :      paul.webster-at-yale.edu
Date: 24 Apr 1996 15:47:14 -0400
Subject: Re: Fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I read the current set of postings about "the best fixative for...", and "the
effect of long term storage on..."and see that many researchers are using PIPES
adjusted to a pH above 7.2 to buffer their fixatives.

This buffer has a pKa of 6.8 so would anyone like to make a comment on the
buffering capacity of this buffer above pH 7.0?

If we have to use a "Good" buffer for fixation, we choose HEPES which has a
higher pKa. We get exceptional morphology in cryosections even if we only use
formaldehyde. We also used it at a conc. of 300mM recently to fix some shark
tissue with similar results. We never did get much from using PIPES.

As for tissue storage, there should be no problem in storing tissue for long
periods in fixative, if the samples are to be examined only for morphology. The
samples will slowly get harder and more brittle the longer they are in
gluteraldehyde but if they are embedded in resin it will not be a problem.
There might be some extraction of soluble material with time but that will only
help with increasing the final contrast of the specimen.

If the samples are only to be fixed in formaldehyde then the fixative should not
be removed. I have heard unconfirmed stories of 10 - 20 year old samples,
stored in formaldehyde at RT, being taken off the shelf and successfully labeled
with antibodies. I guess it all depends on what the antigens were.

Best regards

Paul Webster,
Center for Cell Imaging
Yale University School of Medicine.







From: Glen Macdonald :      glenmac-at-u.washington.edu
Date: Wed, 24 Apr 1996 11:46:30 -0700 (PDT)
Subject: Re: blocking endogenous peroxidase

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Gary,

The question of which solvent to use depends upon what you are trying to
block. The methanol denatures the peroxidase's catalytic site such
that the peroxide cannot be split or released once it has been bound.
Many years ago I compared peroxidase inhibition on bone marrow needle
biopsies between peroxide solutions with PBS,
ethanol, methanol and acetone. PBS and ethanol gave modest
reduction of endogenous peroxidase, and PBS is probably sufficient
for most tissues. Methanol gave the most complete
inhibition of the variety of peroxidatic activities in marrow. Acetone
actually appeared to potentiate it in some leucocytes


Glen MacDonald
Research Scientist
Hearing Research Laboratories of the
Virginia Merrill Bloedel Hearing Research Center
Box 35-7923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156
glenmac-at-u.washington.edu



Gary H. Zajic wrote:

} Hello, I am doing immunocytochemical experiments on mouse cochlea
} using frozen sections. I find that I must block endogenous peroxidase
} in these tissues. Most protocols I see require dehydration to 100%
} EtOH or MeOH when blocking with H2O2. Cochlear tissue is very delicate
} and there is a risk losing some of the fine structure with these
} additional steps. My questions are: why can't I just block with H2O2
} in my buffer? is the alcohol necessary for penetration allowing better
} access of the H2O2 to the tissue? These are 10 micron sections.
} Thanks in advance. Sincerely, Gary
}
}
} zajic-at-umich.edu
}





From: PHOBOS11-at-aol.com
Date: Wed, 24 Apr 1996 21:36:21 -0400
Subject: Re: critical point dryer problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Joyce,

I think that the inlet diaphram maybe clogged. I am now with RMC and still
help people with problems with BAL-TEC / Balzers equipment. Please feel free
to call me at RMC, 520-889-7900. I can talk you through your difficulties if
you call.

P.S I am available to help any BAL-TEC/ Balzers users out there, just E-Mail
me or call.

Best Regards,

Al Coritz




From: ychen-at-MACC.WISC.EDU
Date: Wed, 24 Apr 1996 16:31:32 -0700
Subject: Re: critical point dryer problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} Date: Wed, 24 Apr 1996 11:04:54 -0500
} From: Joyce Craig {Chicago.State.University-at-uxa.ecn.bgu.edu}
} Subject: critical point dryer problems
} To: Microscopy-at-aaem.amc.anl.gov
} X-Sender: bafpjec-at-pop3.ecn.bgu.edu
} X-Mailer: Windows Eudora Light Version 1.5.2
} Mime-Version: 1.0
} Content-Type: text/plain; charset="us-ascii"
}
} We have a Bal-Tec CPD030. It has been used very little. This is because we
} cannot get it to drain. It might take all night for the alcohol-CO2 mixture
} to empty. The instructions seem very straigt-forward. We fill partially
} with 100% ethanol (with or without a sample), cool to 8 degrees C, let in
} the carbon dioxide (yes, the tank is the kind with the dip tube), then press
} the medium out button. At first it seems to empty, then gets very slow.
} Has anyone had a similar problem? The valves seem to work as I hear a loud
} click when pressing buttons. The filters are clean. Any other ideas? thanks.

Joyce,
In my opinion, it looks like the problem is that the exhaust pipe is
jammed. Maybe it is too thin or dirty. If this is the case, in the
beginning it will work fine. After a while, it will freeze and jam. Under
normal conditions, you can hear loud sound and see CO2 gas and ice is
expelled when you open the venting value. If this does not happen, it was
probably jammed.
Ya Chen


Ya Chen

==========================================================================
Cryo/SEM Coordinator
Integrated Microscopy Resource (IMR)-- III M M RRRRRR
an NIH Biomedical Research Resource I M M M M R R
University of Wisconsin, Madison, WI I M M M RRRRRR
1675 Observatory Drive #167 I M M R R
Madison, WI 53706, USA I M M R R
TEL : 608-263-8481 I M M R R
FAX : 608-265-4076 III M M R R
Email:YChen-at-macc.wisc.edu
==========================================================================
IMR WWW Home Page: http://www.bocklabs.wisc.edu/imr.html
2nd Symposium on Integrated Microscopy: Sept. 20-22, 1996






From: Sylvia Francis Zalzal :      franciss-at-ere.umontreal.ca
Date: Wed, 24 Apr 1996 18:33:34 -0400
Subject: Re: critical point dryer problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Date: Wed, 24 Apr 1996 18:28:18
} To: Joyce Craig {Chicago.State.University-at-uxa.ecn.bgu.edu}
} From: Sylvia Francis Zalzal {franciss-at-ere.umontreal.ca}
} Subject: Re: critical point dryer problems
}
} We have in our lab and for over three years the same CPD030 and we never
had this problem. The dehydrant/Co2 exchange should not cause any
trouble,and the sound should not be loud when you depress the buttons.
However we had prior to this Bal-tec 030 we had a cpd020 and a similar
problem occured the reason was " frozen valves " Maybe when cooling the
chamber do not cool under 11 degrees. Try it then call for service check the
valves.
}
} Good luck=20
} Sylvia=20
}
}
}
}
} At 11:04 4/24/96 -0500, you wrote:
} } We have a Bal-Tec CPD030. It has been used very little. This is because=
we
} } cannot get it to drain. It might take all night for the alcohol-CO2=
mixture
} } to empty. The instructions seem very straigt-forward. We fill partially
} } with 100% ethanol (with or without a sample), cool to 8 degrees C, let in
} } the carbon dioxide (yes, the tank is the kind with the dip tube), then=
press
} } the medium out button. At first it seems to empty, then gets very slow. =
=20
} } Has anyone had a similar problem? The valves seem to work as I hear a=
loud
} } click when pressing buttons. The filters are clean. Any other ideas?
thanks.
} }
}
Sylvia Francis Zalzal
Chef de Laboratoire
Laboratoire de Microscopie Electronique
Facult=E9 de m=E9decine dentaire
Universit=E9 de Montr=E9al





From: DENNIS WARD :      dw0005-at-epfl2.epflbalto.org
Date: Thu, 25 Apr 1996 05:09:36 -0400 (EDT)
Subject: Re: ASTM Home Page?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


http://www.astm.org

On 22 Apr 1996, South Bay Technology wrote:

} Can anyone provide me with the URL for the ASTM home page?
}
} Thank you!
}
} Best regards-
}
} David Henriks TEL: 800-728-2233 (toll-free in USA)
} South Bay Technology, Inc. 714-492-2600
} 1120 Via Callejon FAX: 714-492-1499
} San Clemente, CA 92673 USA e-mail: 73531.1344-at-compuserve.com
} sbt-at-msa.microscopy.com
}
} Manufacturers of Precision Sample Preparation Equipment and Supplies for
} Metallography, Crystallography and Electron Microscopy.
}
}




From: Shelley Jan Almburg :      salmburg-at-biology.lsa.umich.edu
Date: Thu, 25 Apr 1996 08:03:40 -0400 (EDT)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604251020.MAA11382-at-elc1.dina.kvl.dk}

please unsubscribe




From: Don Lesher :      72714.265-at-CompuServe.COM
Date: 25 Apr 96 09:03:50 EDT
Subject: ARL SEMQ Microprobe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Any University interested in obtaining an old, but restorable ARL-SEMQ
Microprobe should contact Advanced MicroBeam at 72714.265-at-compuserve.com or
phone: 330 394-1255, fax: 330 394-1834.


Advanced MicroBeam, Inc.
4217C King Graves Rd.
PO Box 610
Vienna, OH 44473 USA





From: Carmine M. Pariante :      cparian-at-emory.edu
Date: Thu, 25 Apr 1996 10:08:26 -0400 (EDT)
Subject: ARL SEMQ Microprobe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


please unsuscribe

Carmine M. Pariante
cparian-at-emory.edu




From: Kurt.Albertine-at-hsc.utah.edu
Date: Thu, 25 Apr 1996 08:51 -0700 (MST)
Subject: Guidance for selling a TEM or two

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello

Jonathan Krupp gave me your eMail address. I have a TEM to sell
and would like to list it on the MSA BB/mail listserver (I am a
member of the MSA). Would you please let me know the format for the
listserver? Also, would you please send instructions for subscribing
to the list via your Web site? Thank you.

Kurt H. Albertine, Ph.D.
Director, Research Microscopy Facility
University of Utah
Health Sciences Center
Salt Lake City, UT 84132

eMail: Kurt.Albertine-at-hsc.utah.edu
FAX: (801) 585-7395
Office: (801) 581-5021




From: Radice-at-urvax.urich.edu (Gary Radice)
Date: Thu, 25 Apr 1996 14:23:15 -0400
Subject: Peltier vs adiabatic CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Alternate-Recipient: prohibited
Auto-Forwarded: prohibited
Content-Return: allowed
Disclose-Recipients: prohibited
Conversion: allowed
Importance: normal
Priority: normal
Sensitivity: Normal

We are looking to buy a new critical point drier to replace our old one
that cooled adiabatically (by venting CO-2) and notice that some CPDs are
now available that cool electronically by the Peltier effect. This seems
like a big advantage but does anyone with experience with both types care
to comment?

Gary Radice 804-289-8107 (voice)
Department of Biology 804-289-8233 (FAX)
University of Richmond
Richmond VA 23173
USA






From: Joyce Craig[SMTP:Chicago.State.University-at-uxa.ecn.bgu.edu]
Date: Thu, 25 Apr 1996 07:42:04 -0400
Subject: critical point dryer problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {01BB3289.A38177E0-at-DENT-BIOMAT2.BU.EDU}
"Microscopy-at-aaem.amc.anl.gov" {Microscopy-at-aaem.amc.anl.gov}

Valve could be freezing. Can you get to it to warm it up?

Ron


----------

We have a Bal-Tec CPD030. It has been used very little. This is because we
cannot get it to drain. It might take all night for the alcohol-CO2 mixture
to empty. The instructions seem very straigt-forward. We fill partially
with 100% ethanol (with or without a sample), cool to 8 degrees C, let in
the carbon dioxide (yes, the tank is the kind with the dip tube), then press
the medium out button. At first it seems to empty, then gets very slow.
Has anyone had a similar problem? The valves seem to work as I hear a loud
click when pressing buttons. The filters are clean. Any other ideas? thanks.








From: rnessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Thu, 25 Apr 1996 09:29:28 -0500 (CDT)
Subject: "Free" Zeiss 10 clarification

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,
I wanted to post more information about the Zeiss TEM that the
University of Iowa's Biology dpartment would like to give away. I spoke
with Shelley Plattner this morning, as he is in charge of the instrument,
and is the person to contact. He can be reached at (319)-335-1070, or
email at shelley-plattner-at-uiowa.edu. He says that it is a 10A, not a 10C.
Zeiss knows that they are trying to give the scope away, and estimates
that it would take about 40 hours of labor -at-$150.00/hr. Zeiss will make
sure that the instrument works upon relocation. Shipping would be extra,
depending on the carrier and distance. Zeiss might be willing to
discount the cost if the person purchases a service contract. You could
discuss this with them by calling 1-800-233-3334. Lam, the Zeiss service
engineer who has maintained this instrument can be reached at extension
725 at the above 800 number. He can answer question in regards to what
would be needed to hook up the microscope.
Thanks to those who have expressed interest, and have tolerated the run
around on our end.

Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu





From: Louisa.Howard-at-Dartmouth.EDU (Louisa Howard)
Date: 25 Apr 96 09:25:50 EDT
Subject: Re: Long Term Fix/Pipes vs. Hepes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {13851929-at-prancer.Dartmouth.EDU}

---Paul Webster wrote:
I read the current set of postings about "the best fixative for...", and "the
effect of long term storage on..."and see that many researchers are using PIPES
adjusted to a pH above 7.2 to buffer their fixatives.

This buffer has a pKa of 6.8 so would anyone like to make a comment on the
buffering capacity of this buffer above pH 7.0?
--- end of quoted material ---

I agree that HEPES would be much better to use at pH 7-3-7.4, given its pKa of
7.4. We use it often in SEM preps. The reason we use PIPES for TEM, is that it
appears to cause less extraction in mammalian and plant tissues. This is
strictly anectdotal on our part, although Coetzee and van Der Merwe, J of
Microscopy135(2) 147-158, 1984, showed that extraction in plant tissue was
greater using HEPES than using PIPES.
I'm definitely willing to try HEPES buffer on mammalian brain tissue, as other
posters have mentioned using HEPES for TEM, with success. Can I use the same
molarity as with PIPES? I currently use a buffer concentraion of 0.1M PIPES pH
-at-7.3, with a gluatraldehyde concentration of 3-4%.
Also, I agree with other posters about the toxicity of Na Cacodylate, although
the HEPES and PIPES buffers, supposedly have carcinogenic problems. If we work
in EM, we have to take all neceesary precautions, when working with these
compounds. I think the PIPES or HEPES buffers might be a better choice for
field work.
thanks.
Louisa Howard
EM facility- Remsen 240
Dartmouth College
Hanover, NH. 03755




From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Thu, 25 Apr 1996 09:32:47 -0500
Subject: Re: "Good" buffers for fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01520d01ada53d735b8d-at-[128.206.15.189]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I must admit I use HEPES buffer, pH 7.2 for my fixatives with great luck
and therefore have not been tempted to change it but I think a good
argument to use PIPES can be made. PIPES has a pKa of 6.8 and HEPES has
one of 7.5. Thus, for a target of pH 7.2, PIPES is 0.4 pH units off and
HEPES is 0.3 units. A good rule of thumb is to use a buffer whose pKa is
within 0.5 units of the desired pH. HEPES may look to be marginally
better than PIPES but fixation solutions tend to acidify during the
fixation step so by having a buffer whose pKa is on the low side of the
starting pH, one would get increased protection at reasonable buffer
concentrations (i.e., the pH would have to drop 0.9 units before being }
0.5 units from PIPES pKa but by only 0.3 units before being } 0.5 units
from HEPES pKa). I think both are "Good" buffers in more than one sense
and both are definitely superior to cacodylate in many ways (e.g., cost,
toxicity, environmental). Furthermore, I believe cacodylate has a pKa of
around 6.3! My feeling has always been the continued prevalence of
cacodylate buffers simply demonstrates that morphologists are some of the
most old-fashioned scientists around and least willing to change. That's
my two cents worth.

} I read the current set of postings about "the best fixative for...", and "the
} effect of long term storage on..."and see that many researchers are using PIPES
} adjusted to a pH above 7.2 to buffer their fixatives.
}
} This buffer has a pKa of 6.8 so would anyone like to make a comment on the
} buffering capacity of this buffer above pH 7.0?
}
} If we have to use a "Good" buffer for fixation, we choose HEPES which has a
} higher pKa. We get exceptional morphology in cryosections even if we only use
} formaldehyde. We also used it at a conc. of 300mM recently to fix some shark
} tissue with similar results. We never did get much from using PIPES.
}
} As for tissue storage, there should be no problem in storing tissue for long
} periods in fixative, if the samples are to be examined only for
} morphology. The
} samples will slowly get harder and more brittle the longer they are in
} gluteraldehyde but if they are embedded in resin it will not be a problem.
} There might be some extraction of soluble material with time but that will only
} help with increasing the final contrast of the specimen.
}
} If the samples are only to be fixed in formaldehyde then the fixative
} should not
} be removed. I have heard unconfirmed stories of 10 - 20 year old samples,
} stored in formaldehyde at RT, being taken off the shelf and successfully
} labeled
} with antibodies. I guess it all depends on what the antigens were.
}
} Best regards
}
} Paul Webster,
} Center for Cell Imaging
} Yale University School of Medicine.

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: Richard Sherburne :      richard.sherburne-at-ualberta.ca
Date: Thu, 25 Apr 1996 15:22:29 -0600
Subject: subscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {01BB32BB.0589A300-at-valerian.mmid.med.ualberta.ca}

subscribe microscopy





From: CSEDAX-at-ARCRIDE.EDU.AR
Date: Thu, 25 Apr 1996 13:10 -0300
Subject: How to face research and service work within the same lab?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi everyone,

since long ago I've been tempted to ask all of you, specially
to those giving TEM/SEM services as the only work activity, your opinion about
the type of services to offer in an institution which is fully oriented to this
activity:


1) How do you consider giving services in the areas of biological materials and
materials science at the same time?

2) Should one focus the work into specific subjects within a given area?

3) How do you define, in advance, the type of work you would be able to do?
or you wait to be asked for a specific task before deciding to do it
or not?

These may sound elemental questions to you but I've never had the chance
to hear people from other systems to give their opinion on that.



Now, for those doing research and giving services...

A) How do you combine both activities?

B) How much time, or effort, do you put on each one of these activities?

C) About the services to offer, should they be only related to the subject
of your research work?


I would appreciate very much any comments and I promise
to make a compilation of all answers for those interested in the subject.
Thanks in advance.

............................................................................
Silvia Montoro
Centro Regional de Investigacion y Desarrollo
Santa Fe
Argentina

csedax-at-arcride.edu.ar





From: p.marks-at-unsw.edu.au (patrick marks)
Date: Fri, 26 Apr 1996 10:40:48 +1000
Subject: How to face research and service work within the same lab?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Netter,

I am working with kangaroo heart muscle, and my aim is to find mitochondrial
volume density and also mitochondrial numbers, but I'm having a problem with
the uranyl acetate and lead citrate staining. I am staining with 4% uranyl
acetate for 1 hour under a lamp, then lead citrate for 4 minutes, yet my
sections are very pale under the transmission microscope, with very little
contrast in many cases. Has anyone else had this problem? I am wondering if
this is a characteristic of this kind of muscle or if it is me and my
methods! If you have any suggestions please RESPOND TO


Dr. SUNIL TEWARI
S.Tewari-at-unsw.edu.au
Patrick Marks
Electron Microscope Unit
University of New South Wales
Sydney 2053
Australia
Email : P.Marks-at-unsw.edu.au





From: p.marks-at-unsw.edu.au (patrick marks)
Date: Fri, 26 Apr 1996 10:45:59 +1000
Subject: Grid staining problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} To: microscopy-at-aaem.amc.anl.gov
} From: p.marks-at-unsw.edu.au (patrick marks)
}
} Hi Netter,
}
} I am working with kangaroo heart muscle, and my aim is to find
mitochondrial volume density and also mitochondrial numbers, but I'm having
a problem with the uranyl acetate and lead citrate staining. I am staining
with 4% uranyl acetate for 1 hour under a lamp, then lead citrate for 4
minutes, yet my sections are very pale under the transmission microscope,
with very little contrast in many cases. Has anyone else had this problem? I
am wondering if this is a characteristic of this kind of muscle or if it is
me and my methods! If you have any suggestions please RESPOND TO
}
}
} Dr. SUNIL TEWARI
} S.Tewari-at-unsw.edu.au
}
Patrick Marks
Electron Microscope Unit
University of New South Wales
Sydney 2053
Australia
Email : P.Marks-at-unsw.edu.au





From: Douglas F Bowling :      al428-at-dayton.wright.edu
Date: Thu, 25 Apr 1996 22:58:32 -0400
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe




From: peling-at-amnh.org (Peling Melville - Interdepartmental Facilities)
Date: Fri, 26 Apr 1996 08:03:34 -0500
Subject: Re: critical point dryer problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510100ada679a5be48-at-[198.116.4.5]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} On Wednesday April 24th, Joyce Craig writes:
}
} } We have a Bal-Tec CPD030. It has been used very little. This is because we
} } cannot get it to drain. It might take all night for the alcohol-CO2 mixture
} } to empty. The instructions seem very straigt-forward. We fill partially
} } with 100% ethanol (with or without a sample), cool to 8 degrees C, let in
} } the carbon dioxide (yes, the tank is the kind with the dip tube), then press
} } the medium out button. At first it seems to empty, then gets very slow.
} } Has anyone had a similar problem? The valves seem to work as I hear a loud
} } click when pressing buttons. The filters are clean. Any other ideas?
} } thanks.
}
}
} Hi Joyce,
}
} We have a Balzers CPD030 and I've experienced the same problem that you
} have. What I have had to do is check the inlet diaphram. 9 times out of
} 10 something has clogged it. I'll take it out and holding the diaphram
} firmly, I "blast" it with an air duster. That usually does the trick.
}
} Peling Melville
}
} --------------------------------------------------------------
} Peling Fong Melville
} Senior Scientific Assistant
} Interdepartmental Facilities
} American Museum of Natural History
} Central Park West at 79th Street
} New York, NY 10024-5192 U.S.A.
} ******************************
} E-mail: peling-at-amnh.org
} Work #: (212) 769-5469
} FAX #: (212) 769-5495

--------------------------------------------------------------
Peling Fong Melville
Senior Scientific Assistant
Interdepartmental Facilities
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 U.S.A.
******************************
E-mail: peling-at-amnh.org
Work #: (212) 769-5469
FAX #: (212) 769-5495






From: brannign-at-asrr.arsusda.gov (Peggy Brannigan)
Date: Fri, 26 Apr 1996 08:46:01 -0400
Subject: Re: low contrast with lead citrate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello microscopists,

Sunil Tewar asked for some advice about lead citrate and low contrast - I
could also use some tips!
My problem is that lead citrate gives variable and unpredictable contrast
in the TEM. I am using LX 112 (plant tissue),osmium, and 4% uranyl acetate
for 45-60 minutes; over the years lead citrate staining times have varied
from 8 min to 25 min. Right now I'm staining around 15 min - was at 18
for a while but suddenly everything was too dense.

I 've read that the pH of the final solution is very important - that it
should be pH 12.0 +/ 1 0.1 When I make it up however the pH is never less
than 12.4! I 've been reluctant to introduce HCl into the picture so I
haven't yet tried reducing the pH, consequently don't know if this is the
problem. I do know that the amount of NaOH in the rinse water is very
important - I use about 5 drops of 1N in 40 mls of water - more than that
bleaches out the tissue. ( Can't remember off the top of my head what that
means in terms of pH). There appears to be no connection between staining
problems and the batch of resin, the type of tissue or the fixation.

One interesting note: when I moved to a new building a year ago, I
couldn't get any staining at all for three months! Non microscopists in
the building were also having unrelated problems, ultimately traced to
impurities in our central distilled water source. Once this was fixed our
problems disappeared.

Has anyone else had problems getting contrast with lead citrate? If so,
how did you fix it?

Thanks, Peggy .

} Hi Netter,
}
} I am working with kangaroo heart muscle, and my aim is to find mitochondrial
} volume density and also mitochondrial numbers, but I'm having a problem with
} the uranyl acetate and lead citrate staining. I am staining with 4% uranyl
} acetate for 1 hour under a lamp, then lead citrate for 4 minutes, yet my
} sections are very pale under the transmission microscope, with very little
} contrast in many cases. Has anyone else had this problem? I am wondering if
} this is a characteristic of this kind of muscle or if it is me and my
} methods! If you have any suggestions please RESPOND TO
}
}
} Dr. SUNIL TEWARI
} S.Tewari-at-unsw.edu.au
} Patrick Marks
} Electron Microscope Unit
} University of New South Wales
} Sydney 2053
} Australia
} Email : P.Marks-at-unsw.edu.au






From: kna101-at-utdallas.edu
Date: Fri, 26 Apr 1996 09:47:03 -0500 (CDT)
Subject: Re: "Good" buffers for fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tom,
I just had to let you know that you made my day. The comment about
microscopists being resistant to change describes my boss to a T. He
recently gave me the task of determining the best way to screen for
mucins in tissue sections, and when I came back, after reading ALOT of
info on mucins, with the suggestion of using lectin staining and
immunohistochemistry (which we have done before for other material), he
pronounced that I would have to work out a technique using ONLY
histochemistry, as he felt the other methods wouldn't work.
Karen P.




From: delphi.beckman.uiuc.edu-at-delphi.beckman.uiuc.edu (steve rogers)
Date: Fri, 26 Apr 1996 10:43:39 -0500
Subject: PIPES for fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: srogers-at-delphi.beckman.uiuc.edu
Message-Id: {v02140b01ada69ecfeda1-at-[128.174.228.241]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} I read the current set of postings about "the best fixative for...", and "the
} effect of long term storage on..."and see that many researchers are using PIPES
} adjusted to a pH above 7.2 to buffer their fixatives.
}
} This buffer has a pKa of 6.8 so would anyone like to make a comment on the
} buffering capacity of this buffer above pH 7.0?

PIPES is one of a few buffers suitable for work with cytoskeletal
components (e.g. microtubules). It is often used to buffer fixativesfor
light & electron microscopy to preserve cytoskeletal structure.


******************************************************
Steve Rogers
Dept. of Cell & Structural Biology and the
Beckman Institute - Optical Visualization Facility
University of Illinois -at- C/U
srogers-at-delphi.beckman.uiuc.edu
******************************************************






From: Charles.P.Daghlian-at-Dartmouth.EDU (Charles P. Daghlian)
Date: 26 Apr 96 12:19:33 EDT
Subject: Sputter coater and parts for JSM T-300

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-id: {13923531-at-prancer.Dartmouth.EDU}

I am helping the Montpelier (Vermont) High School maintain and use a JSM T-300
SEM as part of the science labs. The SEM is linked via its TV output to a Scion
frame grabber in a Macintosh running NIH Image. The science teaching with whom
I am working, Dave McGraw, is involved with the UA Center for Image processing
in Teaching, so this fits in with things that are already being done in the
school. All is well except we need two things.
1. If there is a used T-300 out there that is available for the cost of
shipping, we could use it for spare parts.
2. We are in need of a used sputter coater, too. We have a vacuum pump and can
make some repairs. If you have one available, please let me know.

Thanks in advance to any who respond.

Chuck
*************
Charles P. Daghlian, Ph.D.
Director, Rippel E. M. Facility
Dartmouth College, HB 7605
Hanover, NH 03755
phone 603-646-1039




From: LYNNE (617) 386-1446 :      GARONEL-at-cliffy.polaroid.com
Date: Fri, 26 Apr 1996 14:25:36 -0400 (EDT)
Subject: Open Job Req. at Polaroid for TEM - polymer microscopist

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Polaroid Corp. in Waltham, Ma. USA is looking for a Ph.D. or M.S. in chemistry, materials science, physics or engineering with training in polymer science and
experience in TEM and sample prep. SEM, LM, XPS, XRD, AFM all plusses. STRONG Communication skills and computer skills desirable. Send inquiries to:Lynne
Garone E. Mail:GaroneL-at-Polaroid.com




From: Douglas F Bowling :      al428-at-dayton.wright.edu
Date: Thu, 25 Apr 1996 22:58:32 -0400
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

id {01I3Z4SCD1AO00545V-at-mailhost.wright.edu} for
microscopy-at-sparc5.microscopy.com; Thu, 25 Apr 1996 22:58:33 -0500 (EST)
Received: from selene.wright.edu (selene.wright.edu [130.108.1.20]) by Sparc5.Mi
croscopy.Com (8.6.11/8.6.11) with ESMTP id VAA00352 for {microscopy-at-sparc5.micro
scopy.com} ; Thu, 25 Apr 1996 21:55:46 -0500
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11) id VA
A00355 for dist-Microscopy; Thu, 25 Apr 1996 21:55:47 -0500
Received: from Sparc5.Microscopy.Com (sparc5.microscopy.com [206.69.208.10]) by
ormail.intel.com (8.7.4/8.7.3) with SMTP id VAA14653; Thu, 25 Apr 1996 21:36:24
-0700 (PDT)
Received: from ormail.intel.com by relay.hf.intel.com with smtp
(Smail3.1.28.1 #2) id m0uCfGb-000qDSC; Thu, 25 Apr 96 21:36 PDT




From: kna101-at-utdallas.edu
Date: Fri, 26 Apr 1996 14:44:50 -0500 (CDT)
Subject: Re: "Good" buffers for fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to apologize if I affended anyone on this list with my
"airing" my frustrations. It was inappropriate. Tom's comment just
struck a cord. I realize setting up new protocols for any one tissue is
time consuming and often the "why fix it if it works" attitude is the
best. But, sometimes the old ways don't work as well as you need them
to, then new methods may be necessary.

Karen




From: James S MArtin :      James.S.Martin-at-williams.edu
Date: Fri, 26 Apr 1996 15:32:50 -0400 (EDT)
Subject: microscope hot stages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Seeking ideas for building a low cost microscope hot stage. It doesn't
have to be pretty, only accurate and reliable. We're a small non-profit
museum lab, so I will only ever see a Mettler stage in my dreams. I have
scavenged most of the components necessary for generating and control T
from a PE GC and a melting point apparatus, but am looking for a design
for the stage and insulator jacket (probably water flow).

Any thoughts appreciated. Thanks.

James Martin
Director of Analytical Services and Research
Williamstown Art Conservation Center




From: Stuart McKernan :      mckernan-at-cems.umn.edu
Date: Fri, 26 Apr 1996 10:35:20 -0500
Subject: MMS Spring Symposium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Announcement:

**************************************************************

The 1996 Minnesota Microscopy Society Annual SPRING SYMPOSIUM

**************************************************************

SHERATON INN, MIDWAY, I-94 at HAMLINE AVENUE, ST. PAUL, MN

THURSDAY - MAY 23, 1996

**************************************************************

SCHEDULE OF EVENTS

8:00 - 9:00 AM Coffee and Late Registration

9:00 - 9:45 AM Understanding Video Signals: Cameras, Recording Formats and
Printers
MARTY HARALDSON, Alpha Video, Edina, MN

9:45 - 10:30 AM Coffee Break \ Vendor Displays

10:30 - 11:15 AM Capturing, Storing and Organizing Digital Image Files
MARK SANDERS, Imaging Center, College of Biological Sciences
University of Minnesota, St.Paul Campus

11:15 - 12:45 PM A buffet lunch will be served at the Sheraton Inn

12:45 - 1:00 PM MMS Business Meeting and Election of Officers

1:00 - 1:45 PM Microscopy Resources on the Internet
Dr. STUART MCKERNAN, Center for Interfacial Engineering,
Characterization Facility, University of Minnesota

1:45 - 2:15 PM Coffee Break \ Vendor Displays

2:15 - 3:00 PM Public Domain Shareware, Microscopy and the Internet
Dr. DAVID BRIGHT, 1996 MSA Traveling Speaker
Research Chemist, Microanalysis Group
NIST, Gaithersburg, MD

3:00 - 4:30 PM Vendor Displays

*********************************************************************
Please make your reservation in advance! NO LATER THAN MONDAY, MAY 20, if you
plan to attend the Symposium (preferred, or pay at the door).

Contact Stuart McKernan at (612) 626-7942, stuartm-at-maroon.tc.umn.edu or
Dwight Erickson at (612) 736-2830, usmmm214-at-ibmmail.com

Symposium Fee:
$20.00 current regular MMS members 95/96, $10.00 student members 95/96
$30 non-member(confers regular membership),
$15.00 non-member students (confers student membership).

*********************************************************************

Vendors: A number of vendors of digital and other microscopy equipment will be
present at the Symposium. If your company would like to have table space for
product display, please contact Symposium Vendor Liaison: Diana Kittleson at
Pillsbury TPC Labs, 330 University Ave. S. E., Minneapolis, MN 55414,
Phone: (612)330-1898 Fax: (612)330-8266

**********************************************************************


Stuart McKernan stuartm-at-maroon.tc.umn.edu
CIE Microscopy Center, University of Minnesota Office: (612) 626-7942
100 Union Street S. E., Minneapolis, MN 55455 Lab: (612) 624-6590





From: beebed-at-alize.ERE.UMontreal.CA (Dwight Beebe)
Date: Fri, 26 Apr 1996 11:47:10 -0500
Subject: Re: "Good" buffers for fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01510102ada6adb905ba-at-DialupEudora}
Mime-Version: 1.0
Content-Type: text/plain; charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

Tom Phillips said:
} I must admit I use HEPES buffer, pH 7.2 for my fixatives with great luck
} and therefore have not been tempted to change it but I think a good
} argument to use PIPES can be made.

Hi,
I'd like to chime in on this thread. I have been using PIPES as a
buffer, but only in fixations for immunolocalizations. I had used it for
morphological studies, but following a conversation with MV Parthasarathy
(EM Director, Cornell), I switched back to either phosphate or cacodylate.
My original interest stems from the following report: Salema R, I Brandao
1973 The use of PIPES buffer in the fixation of plant cells for electron
microscopy. J Submicr Cytol 5:79-96. As I work exclusively with plant
material, primarily leaves, it appeared to be best. The paper presents
results indicatng that PIPES is the least extractive of buffers currently
in use at the time. However, the work by Coetzee and van der Merwe (Coetzee
J, CF van der Merwe 1987 Some characteristics of the buffer vehicle in
glutaraldehyde-based fixatives. J Microsc 146:143-155.) indicates that the
least extraction was found with phosphate-buffered glutaraldehyde. Again,
their work was conducted using plant material (bean leaves) and may not be
directly applicable to animal tissue, which was the subject of the original
question. As most people know, phosphate is not compatible with added
salts, hence, cacodylate would then be the buffer of choice. I would agree
that morphologists are conservative in their approach to new techniques. I
was extremely sceptical of the use of microwaves in fixations, until I was
able to use one of the units myself. I am now convinced that this is a
very powerful technique and yields fixation images equal to or superior to
conventional protocols. The time-saving aspect is a tremendous bonus! I
was loaned a Pelco unit by Phillip Slakmon of Scott Scientific for a number
of months and have used it for more than a dozen fixations, which included
standard protocols for both morphological and immunocytochemical studies.
I have polymerized Spurr's resin, Epon-Araldite, LR White, and mixtures of
Spurr and Epon, all with success (good infiltration, polymerization,
sectioning, staining, and stability in the beam as criteria). Note that
the polymerizations were conducted under water (!) in BEEM capsules. Has
to be tried to be believed, I guess.
Anyway, I would hope that Partha would add his comments to this
discussion on buffers, particularly regarding the suitability of PIPES for
morphological work.


Dwight U. Beebe E-mail: beebed-at-ere.umontreal=
.ca
Institut de recherche en biologie v=E9g=E9tale Voice: 514-872-4563
Universit=E9 de Montr=E9al FAX: 514-872-9406
4101, rue Sherbrooke est
Montr=E9al, Qu=E9bec H1X 2B2
Canada






From: Joyce Craig :      Chicago.State.University-at-uxa.ecn.bgu.edu
Date: Fri, 26 Apr 1996 16:34:40 -0500
Subject: Halobacterium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604262237.RAA01018-at-uxa.ecn.bgu.edu}
X-Sender: bafpjec-at-pop3.ecn.bgu.edu (Unverified)
X-Mailer: Windows Eudora Light Version 1.5.2
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Is anyone out there working with halobacteria? Does anyone have pictures?





From: chandler-at-lamar.ColoState.EDU (John Chandler)
Date: Fri, 26 Apr 1996 17:37:25 -0600
Subject: Looking for polarizing stereo viewers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We need to find a supplier of stereo viewers, the cardboard-mounted
polarizing kind. I've looked in every EM supply house catalog that I have
and can't find them anymore. I'd appreciate any help in finding these.
Thanks.

John
chandler-at-lamar.ColoState.EDU






From: James S MArtin :      James.S.Martin-at-williams.edu
Date: Fri, 26 Apr 1996 22:34:28 -0400 (EDT)
Subject: for microscope manufacturers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Several months ago I posted a request for microscope manufacturers to
contact me with information on costs for epi-fluorescence upgrades to
polarizing light microscopes. The information will be presented at a
national conference.

So far, I have information only from Olympus. Anybody else like to be
represented?

James Martin
Director of Analytical Services and Research
Williamstown Art Conservation Center
Williamstown, MA




From: cmurphy-at-GGPL.ARSUSDA.GOV
Date: Fri, 26 Apr 1996 14:02:24 -0400
Subject: frozen osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings fellow microscopist: Our usually dependable refrigerator turned on us one day recently. It dropped to 0C and our 4%
aqueous osmium tetroxide froze. Does anyone know if this is still good to use. Please drop me a line if anyone has had similar
experiences. Thanks Charlie Murphy




From: Mike Folsom :      mwfolsom-at-unm.edu
Date: Sat, 27 Apr 1996 15:08:57 -0600 (MDT)
Subject: Re: frozen osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To: microscopy-at-Sparc5.Microscopy.Com

On Fri, 26 Apr 1996 cmurphy-at-GGPL.ARSUSDA.GOV wrote:

} Greetings fellow microscopist: Our usually dependable refrigerator turned on us one day recently. It dropped to 0C and our 4%
} aqueous osmium tetroxide froze. Does anyone know if this is still good to use. Please drop me a line if anyone has had similar
} experiences. Thanks Charlie Murphy
}

Yes -

I usually keep my osmium solutions frozen. They appear to work just fine -

Michael






From: rgwhite-at-vaxc.cc.monash.edu.au (Rosemary White)
Date: Sun, 28 Apr 1996 14:31:08 +1200
Subject: Re: low contrast with lead citrate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have also had this problem intermittently. It's almost always been
solved by making up new lead citrate, being scrupulous about purity of
distilled water, following the recipe exactly, and making sure the pH is
above 11 (usually ends up higher than this anyway - but may not stay that
high). There are various lead citrate recipes, once we had to switch to
the original Reynolds recipe for some reason, after a while someone tried
the usual recipe and it worked again.... Staining times have varied too,
usually it depends on how old the lead solution is - increasing time for
older solutions.

With plant and animal tissue, we use alcoholic U acetate for only 10-15 min
(made fresh from 4% aqueous - add equal quantities of 4% UA, distilled
water and 100% ethanol, leave for 10 min till no more bubbles), and no NaOH
in rinse water - we just warm it slightly - too warm and sections lose
contrast.

good luck!


Rosemary White
__ /
Department of Ecology _/ \__/ \
and Evolutionary Biology / \
Monash University / Australia \
Clayton, Victoria 3168 \ ____ /
phone 61-3-9905 5670 \_/ \_*_/
fax 61-3-9905 5613 __
email rgwhite-at-vaxc.cc.monash.edu.au \/






From: Vladimir.Oleshko :      oleshko-at-uia.ua.ac.be
Date: Sun, 28 Apr 1996 15:48:30 +0200 (MET DST)
Subject: Re: Open Job Req. at Polaroid for TEM - polymer microscopist

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




On Fri, 26 Apr 1996, LYNNE (617) 386-1446 wrote:

} Polaroid Corp. in Waltham, Ma. USA is looking for a Ph.D. or M.S. in
chemistry, materials science, physics or engineering with training in
polymer science and=20
} experience in TEM and sample prep. SEM, LM, XPS,
XRD, AFM all plusses. STRONG Communication skills and computer skills
desirable. Send inquiries to:Lynne=20
} Garone E. Mail:GaroneL-at-Polaroid.com
} =20

Dear Sir,

I would like to express my interest in your proposal. Enclosed is my=20
resume. As it indicates, I have a solid physicochemical background, both=20
from my education at the Moscow University and my subsequent professional=
=20
experience including training in polymer science.

I am currently working as a visiting scientist at the Department of=20
Chemistry, University of Antwerp (U.I.A.), Belgium in close collaboration=
=20
with Prof. Dr. Renaat Gijbels and Prof. Dr. Willem Jacob. At this=20
department well as earlier at the Photochemistry Department of the=20
N.N.Semenov Institute of Chemical Physics, Russian Academy of Sciences
I am primarily involved in analytical electron microscopy studies of=20
silver halide photographic systems, colloids and small particles.

I am familiar with the following methods and techniques: conventional and=
=20
energy-filtering TEM, SEM, STEM, EDX, EELS, image analysis, IR-,=20
UV-VIS-spectrometry, Raman-spectroscopy, electron spin resonance,=20
scanning tunneling microscopy, working with PC.

I am deeply impressed by your proposal. I am hard working, and I feel=20
strongly that my motivation and enthusiasm could be of benefit to your=20
programs.

I look forward to hearing from you.

Sincerely Yours,

Vladimir Oleshko
=20
Curriculum Vitae of Vladimir P. Oleshko, Ph.D.
Personal data
Date of birth: September 19, 1953.
Place of birth: Kobishcha, Bobrovitsky district,
Chernigovsky region, Ukraine.
Martial status: married with son and daughter.
Fluent in Russian and in English.

Permanent address: Photochemistry Department, N.N. Semenov Institute of
Chemical Physics, Russian Academy of Sciences, Novatorov Str. 7a, 117421
Moscow, Russia. =20
Tel.: 7 (095) 9363950 (office); Fax: 7 (095) 936 3950;=20
Tel.: 7 (095) 5504730 (home).=20

Current address: Micro- and Trace Analys is Centre (MiTAC), Chemistry
Department, University of Antwerp (U.I.A.), Universiteitslpein 1, B-2610
Antwerpen-Wilrijk, Belgium. Tel.: +32-3-820.23.64; Fax: +32-3-820.23.76;=20
E-mail: oleshko-at-uia.ua.ac.be=20

Education=20
1974 B.S. (Chemistry) Moscow State University, Moscow, USSR. Honours in
Chemistry.=20

1976 M.S. (Physical Chemistry) Moscow State University, Moscow, USSR.=20
Honours in Physical Chemistry.=20

1983 Ph.D. (Chemistry) Moscow State University, Moscow, USSR.=20

Ph.D. Thesis "Application of spin trapping to investigation of
adsorption and catalytic processes".

Professional Experience
1971-1976: Student, Chemistry Department, Moscow State University.=20
Extended university courses and practical training in inorganic,
analytical, organic and physical chemistry, polymer science and chemical
technology and additional extended courses in high mathematics, physics
and physical chemistry (catalysis and chemical kinetics, adsorption
phenomena, electrochemistry, physical chemistry of solid state, quantum
chemistry, and physical instrumental methods in chemistry).=20

1976-1982: Postgraduate student, N. I. Kobozev Laboratory of Catalysis
and Gas Electrochemistry, Physical Chemistry Division, Chemistry
Department, Moscow State University.=20
Experience in physical chemistry, catalysis, adsorption phenomena,
kinetics of chemical reactions, vacuum and adsorption techniques and
characterization of oxide based catalytic system s by electron spin
resonance using spin probe and spin trap techniques, infrared-, and
electron spect rometry in the ultraviolet and visible regions and computer
simulation of donor-acceptor interactions on the oxide surfaces.=20

1982-1984: Junior Research Fellow, Research Institute of Chemical
Technology, Moscow region, USSR.=20
Experience in polymer science of composite polymer systems and
characterization of polymer molecular dynamics and stability by spin probe
technique.=20

1984-1988: Head of research group, Senior Research Fellow, Experimental
Designing Bureau "Horizon", Moscow, USSR. =20
Experience in materials science and characterization of superconductors
(Nb-Ti alloys, Nb3Sn, and high-temperature superconductor Y-Ba-Cu-O
systems) by scanning (SEM) and transmission electron microscopy (TEM) and
energy-dispersive X-ray microanalysis (EDX).=20

1988-1993: He ad of research group, Senior Research Fellow,
Photochemistry Department, N.N.Semenov Institute of Chemical Physics,
Russian Academy of Sciences, Moscow, Russia. =20
Experience in imaging science and chemical physics of record media (silver
halide photographic systems, thin films, small particles and nanoclusters,
photosensitive semiconductor dispersions and their characterization by
analytical electron microscopy (AEM) methods (TEM/SEM/STEM/electron
diffraction/EDX), image analysis and scanning tunneling microscopy.=20

1992.08.-11.: Visiting Scientist, Korea Research Institute of Chemical
Technology, Daedeog-dangi, Taejeon, South Korea. =20
Experience in photographic science and technology of silver halide
emulsions.=20

1993.08.-to present: Visiting Scientist, Micro- and Trace Analysis Centre
(MiTAC), Chemistry Department, University of Antwerp (U.I.A.),
Antwerpen-Wilrijk, Belgium. =20
Experience in energy-filtering transmission electron microscopy (EFTEM)
and electron energy-loss spectroscopy (EELS) and scanning
energy-dispersive X-ray microanalysis (STEM/EDX) of disperse many-particle
systems (silver-halide-based photographic systems, metal and metal sulfide
colloids, polynuclear cluster coordination compounds, nanocrystalline
films) and Monte Carlo simulations of electron beam-solid interactions.=20

Number of papers: 51 scientific publications.

Invited reviews: Characterization of complex silver halide photographic
systems by means of analytical electron microscopy //Microbeam Analysis.
1995. V.4. N.1, pp. 1-29 (first author).=20

Scanning Microanalysis: In Handbook of Microscopy. Eds. by S. Amelinckx et
al., VCH, 39 p. 14 ill. (1996) (submitted, first author).

Presentations:=20
20 presentations at international and national scientific congresses and
conferences.=20

Research interests
1. Surface phenomena, adsorption, photochemistry, photophysics and catalysi=
s=20
at surfaces of solids, nanochemistry and nanophysics of molecular organized=
=20
systems.

2. Analytical electron microscopy (energy-filtering TEM/ESD/EELS,
STEM/SEM/EDX), image analysis and scanning tunneling microscopy of
disperse matter.=20

3. Formation of ultradisperse metal and metal sulfide phases on
microcrystals of silver halide emulsions at different stages of the
photographic process i.e., ripening, exposure, and development.=20

4. Cluster formation, aggregation and transformations in disperse
many-particle-systems (silver halide emulsions, polynuclear cluster
coordination compounds, colloids, nanocrystals, semiconductor dispersions,
mechanically activated high-temperature superconductors).=20

Main professional accomplishments=20
1. Elucidation of mechanisms of acid-base interactions on oxide surfaces
during radical catalytic reaction by spin traps and spin probes.=20

2. Development of a technique for evaluation of a surface oxide bacidity
by the probing reaction of catalytic radical decomposition of
2-methyl-2-nitrosopropane.

3. Development of a technique for estimation of fine structure, core,
shell and outer sizes of drops of color coupler dispersions in color
photograph ic materials by combined positive-negative staining of
ultrathin sections following stereological reconstruction of volume sizes.=
=20

4. Elucidation of mechanisms of selective and non-selective reduction of
silver halide emulsions with a series of developers (black-and-white,
color, volume and surface developers) and with NaBH4. Demonstration of
formation of percolation networks of silver filaments and particles in the
course of development with an active volume developer.

5. Demonstration of fractal aggregation of products of chemical
sensitization on tabular microcrystals of silver halide emulsions in the
course of the ripening.=20

6. Development of the optimal strategy of structural and analytical
characterization of complex silver halide photographic systems by the
combination of AEM and image analysis techniques.=20

7. Evaluation of crystalline and electronic structure and elemental
composition of advanced double structure tabular microcrystals of silver
halide emulsions by the combination of cryo-EFTEM/ESD/EELS and
cryo-STEM/EDX techniques. Development of a modified log-ratio EELS
technique for estimation of the local crystal thickness.=20

Professional Societies=20
Russian D. I. Mendeleev Chemical Society=20
Belgian Society for Microscopy
European Microanalysis Society

References=20
Prof. Dr. Renaat H. Gijbels=20
Micro- and Trace Analysis Centre (MiTAC), Co-Director
Department of Chemistry, University of Antwerp (U.I.A.),
Universiteitsplein, 1, B-2610 Antwerpen-Wilrijk, Belgium.=20
Tel.: +32-3-820.23.60 Fax: +32-3-820.23.76

Prof. Dr. Willem A. Jacob=20
Electron Microscopy Centre, Head
Department of Medicine, University of Antwerp (U.I.A.),
Universiteitsplein, 1, B-2610 Antwerpen-Wilrijk, Belgium.
Tel.: +32-3-820.25.09 Fax: +32-3-820.26.03=20

Prof. Dr. Gustaaf Van Tendeloo
EMAT, University of Antwerp (R.U.C.A.),=20
Groenenborgerlaan 171, B-2020
Antwerp, Belgium.
Tel.: +32-3-218.02.62 Fax: +32-3-218.02.57.

Prof. Dr. Valery V. Lunin
Department of Chemistry, Dean
Moscow State University,
Vorob=92evy Gory, 119899 Ms cow, Russia
Tel.: 7 (095) 939 45 75, 7 (095) 939 35 71
Fax: 7 (095) 932 00 67.

Prof. Dr. Michael V. Alfimov
Department of Photochemistry, Director
N.N. Semenov Institute of Chemical Physics,
Russian Academy of Sciences Novatorov Str., 7a,=20
117421 Moscow , Russia.
Tel: 7(095) 936 77 53 Fax: 7(095) 936 12 55.





From: Stephen Edgar :      s.edgar-at-auckland.ac.nz
Date: Mon, 29 Apr 1996 10:13:03 NZS
Subject: Re: venting print processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



The following book may be of interest in regard to this discussion
thread:

"Health Hazards for Photographers" by Siegfried & Wolfgang Rempel.
Published by Lyonn (? may be an abbreviation) in 1993.
ISBN 1558211810

I have not seen this text myself as it is too expensive in this part
of the world - only $16-95 in the US though.

Perhaps Donna Wagahoff's building engineer was thinking of X-ray
processors rather than ordinary photographic machines when he
expressed concern over the extraction of fumes. I have no experience
of X-ray processors myself, but understand that they have caused
problems for a lot of users - isn't formaldehyde used as a hardener
in these machines?


Regards

Stephen Edgar

Electron Microscope Unit, Pathology Department
School of Medicine
University of Auckland
Private Bag 92019
Auckland
New Zealand

email address: s.edgar-at-auckland.ac.nz
Phone : +64-9-3737599 extn 6473 (GMT + 12h)
Fax : +64-9-3737459




From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Sun, 28 Apr 1996 10:45:51 -0600
Subject: Re: Looking for polarizing stereo viewers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130500ada94fc6a1fa-at-[131.230.97.69]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

John Chandler posed the following question:

} We need to find a supplier of stereo viewers, the cardboard-mounted
} polarizing kind. I've looked in every EM supply house catalog that I have
} and can't find them anymore. I'd appreciate any help in finding these.
} Thanks.
}

One possible solution:

Reel 3-d Enterprises Inc...Po Box 2368...Culver City, CA 90231-2368
Phone: (310)837-2368


I just posed a similar question for red/greed glasses. Several kind
respondants gave me the name of the above company which sells all kind of
3D materials - including the polarizing glasses. I received the r/g glasses
the next day. They accept charge cards or checks only (no PO's, etc).
Prices are so reasonable, that I just ordered them on my own charge card.

I am just a happy patron with no financial interests in the company.

Peace -



#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: Brett W. Cockman :      bcockman-at-uniwa.uwa.edu.au
Date: Mon, 29 Apr 1996 10:03:32 +0700 (WAST)
Subject: RE: frozen osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604290027.TAA02614-at-Sparc5.Microscopy.Com}

Hello Charlie,
I store my aqueous osmium solutions (usually 4%) in a frozen state and have
had no problems to date. It is a convenient way to keep the osmium over a
longer period as I only need to use it occasionally and saves the fridge
interior from being inadvertently redecorated with the 'black' precipitate.
Regards,
.
.
Brett W. Cockman
Technologist in Charge
School of Dentistry
University of Western Australia
Voice: (619-2205834)
Fax: (619-2213829)
e-mail; bcockman-at-uniwa.uwa.edu.au




From: GeoffA :      geoffa-at-amsg.austmus.oz.au
Date: 24/4/96 8:59 AM
Subject: long term storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To expand upon Louisa Howard's question of long term storage -she is
asking about storage in cacodylate buffer.

The question has recently been asked here about long term storage in
phosphate buffers. We have a variety of fish, rat and cow tissues that were
collected (or will be collected) "just in case" EM may be needed. Some of
these tissues have been around for at least 10 years and will likely be here
another 10 years.

What are the current thoughts on storage for long lengths of time?
Would it be better to keep the tissues in fixative as opposed to buffer?
Would cacodylate buffer be preferable to phosphate (or is there something
else)? Any special techniques for processing old tissues and getting useable
results? How long have tissues been kept around and still been successfully
worked with?

Thanks for any advice from all the experts on the list.

Rosemary Harris
Electron Microscopy Laboratory
Alberta Environmental Centre





From: em-at-mediacity.com (Ed Monberg)
Date: Sun, 28 Apr 1996 23:32:32 -0800
Subject: Re: venting print processors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01540b01adaa1e640d84-at-[205.216.172.10]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} The following book may be of interest in regard to this discussion
} thread: etc.


Gee, guys, doesn't anyone read the labels?

. . and take a browse in the stacks in "toxicology" ?


(Can it be much more complex,

but a lot more dangerous I suspect, than that,

When you can later "taste" the chemicals wetting your skin,

as I have experienced myself.) Sorry about the sentence,

but those chemicals are getting to my brain.



Regards,



(signed) Ed Monberg {em-at-mediacity.com}

--------------------------------------------------

510-429-1060 Fax 429-1065
LMDC, (Laser Motion Development Co.)
3101 Whipple Road
Union City, CA 94587-1216


For our Most recent Catalogue of "On Hand" EQUIPMENT:
Send empty mail to: {Cat-at-lasermotion.com}


Our web page: http://www.lasermotion.com (Is beginning to take shape!)
Our e-mail: office-at-lasermotion.com

{-------------------------------- Our page width
-----------------------------}






From: Koenraad.Janssens-at-mtm.kuleuven.ac.be (Koenraad Janssens)
Date: Mon, 29 Apr 1996 09:35:34 +0200
Subject: SIMCON software, shareware?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Fellow Microscopists,

The last few years I have been involved in research on localized (near
nanometer resolution) strain characterization using two-beam electron
diffraction contrast imaging (an operational mode of transmission
electron microscopy). See any book on transmission electron microscopy
for the basics.

Part of this project was to develop a software package allowing for the
analysis of strain fields of arbitrary geometry. This software allows
one to model a microscopic strain field with various mathematical methods
(including finite elements) and subsequently verify this model by
comparison with experimental observations in the TEM.

At this moment we would like to know whether anybody would be interested
in this software (SIMCON, i.e. SIMulating CONtrast images). If there are
enough people interested we will probably arrange for a shareware
software release. If you are indeed one of the interested please let me
know by e-mail. To keep track of SIMCON you might check on
http://www.mtm.kuleuven.ac.be/~janssens/simcon.html

Friendly Greetings,

Koen Janssens

=============================================================

Koenraad G. F. Janssens, Dr. Ir.
Katholieke Universiteit Leuven (KUL)
Departement Metaalkunde en Toegepaste Materiaalkunde (MTM)
de Croylaan 2, B-3001 Leuven
Belgium
Tel: +32-16-32.1232
Fax: +32-16-32.1992
Koenraad.Janssens-at-mtm.kuleuven.ac.be




From: MR A HALL, Elektronmikroskopie, X3297 :      HALL-at-scientia.up.ac.za
Date: Mon, 29 Apr 1996 11:56:06 GMT+2
Subject: Buffers,Fixatives,etc

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Microscopists

With regard to all the dicussion on fixatives, buffers,etc; maybe
it's time to suggest two references:

1.Coetzee,J.and van der Merwe, C.F.,Some characteristics of the
buffer
vehicle in gluteraldehyde-based fixatives.Journal of Microscopy,Vol
146 Pt2, May 1987,pp143-155.

2.Coetzee,J. and van der Merwe, C.F.,The influence of processing
protocol on the ultrastructure of bean leave cells.S.African Journal
of Botany,1986 52(2)

Alan Hall
Unit for Electron Microscopy
University of Pretoria
Pretoria.
Tel: +27+012-420 3297
Fax: +27+012-420 3266




From: Vladimir.Oleshko :      oleshko-at-uia.ua.ac.be
Date: Mon, 29 Apr 1996 12:32:36 +0200 (MET DST)
Subject: Apology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Please accept my apologies for the not intentional mistake. My message
was addressed just to the person posting the job announcement.

Vladimir Oleshko




From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Mon, 29 Apr 1996 09:31:52 GMT
Subject: FYI

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} } __________________________________________________________________________
_____
}
} } } } HERE IS SOMETHING THAT WE ALL NEED TO TAKE ACTION ON. PASS THIS ON
} } } } TO YOUR FRIENDS.
} } } }
} } } } } THE NEO-NAZIs are trying to start a news group. Their purpose is to
} } } } get
} } } } } their message of hate out to young people using the Internet. Its
} } } } } originator, Mr. Burdi, is confident they will get the "yes" votes they
} } } } need.
} } } } } He said: "Let me be perfectly blunt and state that we have more than
} } } } enough
} } } } } 'net-nazis' to win this thing hands-down. But every one of you must
} } } } vote
} } } } } 'yes'".
} } } } }
} } } } } Mr. Burdi's confidence is disturbing, so please give this letter the
} } } } widest
} } } } } possible distribution and help us defeat this news group with a your
} } } } "No"
} } } } } vote.
} } } } }
} } } } } News groups are public discussion areas of the Internet and their
} } } } formation
} } } } } requires enough support from the Internet community. EACH AND
} } } } EVERYONE OF
} } } } US
} } } } } HAS ONE VOTE when it comes to creating a new Usenet group. Please
} } } } vote
} } } } } against this news group.
} } } } }
} } } } } In order to vote, E-mail: music-vote-at-sub-rose.com. Your mail message
} } } } should
} } } } } contain only one statement: I vote NO on rec.music.white-power.
} } } } }
} } } } } Vote counting is automatic. Failure to follow these instruction may
} } } } mean
} } } } } that your vote does not get counted. If you do not receive an
} } } } acknowledgment
} } } } } of your vote within three days, contact the vote taker about the
} } } } problem.
} } } } }
} } } } } Please distribute this message.
}
} Dr. Sheldon M. Schuster
} Professor and Program Director
} Box 110580
} University of Florida
} Gainesville, FL 32611-0580
} Phone -- 352-392-8408
} Fax -- 352-392-8598
}
}
}
*******************************************************
Greg Erdos Phone: 352-392-1295
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611 http://www.biotech.ufl.edu/~emcl/

*******************************************************





From: Fermin, Cesar :      fermin-at-TMC.Tulane.Edu
Date: 29 Apr 1996 10:23:18 -0500
Subject: Montage FR1 slide

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Slide making machine:
We have a Montage FR1 slide making machine that worked fine for us in older
Macs with systems older than 7.5.2. My secretary told me that the company
went out of business and that the phone was disconnected.

1) Does anyone knows if the compony was bought?
2) Has anyone newer drivers that possible works on newer macs?

I would appreciate any information you can provide and solution. As now, I
contemplate hooking it up to an older mac.

*********************************************************************
*Cesar D. Fermin, Ph.D \|T|/ Fax (504) 587-7389 *
*Tulane Medical School /|M|\ Voice Mail (504) 584-2618 *
*Pathology/SL79 \|C|/ Secretary (504) 584-2436 *
*New Orleans La 70112-2699 /|*|\ Lab/Techn. (504) 584-2521 *
* {Fermin-at-tmc.tulane.edu} ----} Prof. Pathology & Otolaryngology *
*http://www1.omi.tulane.edu/departments/pathology/fermin/cdftop.html*
*********************************************************************




From: William R Oliver :      oliver-at-ipas4.afip.mil
Date: Mon, 29 Apr 1996 11:40:21 -0400 (EDT)
Subject: Re: FYI

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Why is this on a microscopy newsgroup? If we are all going
to start posting about our particular political hobby horses,
just let me know. I got a million of 'em.


billo





From: jofu-at-enh.nist.gov (Joe Fu)
Date: Mon, 29 Apr 1996 11:45:25 -0400
Subject: Re: FYI

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604291704.NAA29393-at-leatherback.nist.gov}
X-Sender: jofu-at-mailserver.nist.gov
X-Mailer: Windows Eudora Version 2.0.3
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Greeting:


Does anyone know the OSRAM Inc. telephone# & address? We need bulb
replacement for our microscope.
Yhank you.

Joseph Fu
NIST Rm.A117 Bldg.220
Gaithersburg, Md. 20899
Tel:301-975-3495
e mail: jofu-at-enh.nist.gov
Fax: 301-869-0822





From: Ann-Fook Yang :      YANGA-at-em.agr.ca
Date: Mon, 29 Apr 1996 12:02:44 -0400
Subject: Frozen osmium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {s184b0cb.096-at-EM.AGR.CA}
X-Mailer: Novell GroupWise 4.1

Charlie was worrying whether frozen osmium tetraoxide was still
good for fixation. Michael and Brett have their osmium stored
frozen.
A colleaque in another lab. liked to store 4% aqueous oxmium in
the freezing compartment. She had had broken ampules on several
occasions. This came to light only after her retirement when her
successor asked me how I stored mine.
I make aqueous solution, seal 0.5ml in each ampule and leave them
in the fume hood; no more osmium fume in the fridge and no black
cooling compartment to worry about.

Ann Fook Yang
Yanga-at-em.agr.ca






From: Jay Jerome :      jjerome-at-bgsm.edu
Date: Mon, 29 Apr 1996 12:38:45 -0400 (EDT)
Subject: Re: Looking for polarizing stereo viewers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John et al-

Although no longer listed in the catalog, Ted Pella still carries
polarizing glasses for stereo viewing- or at least did a month ago.


Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-bgsm.edu *
**************************************************************

On Fri, 26 Apr 1996, John Chandler wrote:

} We need to find a supplier of stereo viewers, the cardboard-mounted
} polarizing kind. I've looked in every EM supply house catalog that I have
} and can't find them anymore. I'd appreciate any help in finding these.
} Thanks.
}
} John
} chandler-at-lamar.ColoState.EDU
}
}
}




From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Mon, 29 Apr 1996 09:52:01 GMT
Subject: Re: Question: digital image recording for metallograph

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 10:16 AM 4/25/96 EDT, you wrote:
}
} Our lab would like to convert our Zeiss Axiomat from film to digital.
} We have no direct experience, and would appreciate any comments or
} experiences that would help us research the best route to take, both
} hardware and software, but particularly the former.
}
} I believe that this topic has been discussed recently, but unfortunately
} I was not concerned until just now! If anyone has saved either their own
} previous postings or those of others, I would appreciate if you would
} email to me directly if you consider it repetitious to repost.



Hi Pat,
I maintain an archive of most of the biological and computer
related questions and responses that appear in this list. If you go to the
www address at the end of this page and click on the "Tips & Tricks Wizard"
you should find a discussion called "Acquiring digital images". Let me know
if you do not have www access and I will be hapy to E-mail it to you.




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {

Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 904-392-1295
ICBR EM Core Lab fax 904-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/





From: ckblack-at-dow.com
Date: Mon, 29 Apr 1996 11:13:51 -0400
Subject: Re: FYI

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





From: ckblack-at-dow.com
Date: Mon, 29 Apr 1996 11:13:51 -0400
Subject: Re: FYI

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In response to....



} } } } HERE IS SOMETHING THAT WE ALL NEED TO TAKE
ACTION ON. PASS THIS ON
} } } } TO YOUR FRIENDS.
} } } }
} } } } } THE NEO-NAZIs are trying to start a news group.
Their purpose is to
} } } } get
} } } } } their message of hate out to young people using
the Internet.........etc.

------------------------
This forum (microscopy) is for the discussion of
microscopy related material. Political issues and so
forth have NO place here. This is a place where
scientists can discuss science.

We all have opinions concerning issues of censorship
and other such controversial things concerning the
Internet. There are appropriate forums for
addressing your concerns......but not here.


Respectfully.

Cary Black
Dow Chemical

Greg Erdos
Phone: 352-
392-1295
Scientific Director,


ICBR Electron Microscopy Core Lab

218 Carr Hall
Fax: 352-846-
0251
University of Florida E-mail:
gwe-at-biotech.ufl.edu
Gainesville, FL 32611
http://www.biotech.ufl.edu/~emcl/

******************************************************
*

================== RFC 822 Headers ==================



From: gwe-at-biotech.ufl.edu (Greg Erdos)
Date: Mon, 29 Apr 1996 09:31:52 GMT
Subject: FYI

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
Mon, 29 Apr 1996 10:41:29 -0400
Received: from sparc5.microscopy.com by
inet2.hs.dow.com with SMTP id AA15161
(InterLock SMTP Gateway 3.0); Mon, 29 Apr 1996
09:37:56 -0500
Received: (from daemon-at-localhost) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) id IAA03660 for
dist-Microscopy; Mon, 29 Apr 1996 08:23:57 -0500
Received: from snitch.biotech.ufl.edu
(snitch.biotech.ufl.edu [128.227.186.64]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id
IAA03657 for {Microscopy-at-sparc5.microscopy.com} ; Mon,
29 Apr 1996 08:23:55 -0500
Received: from empc1.biotech.ufl.edu by
snitch.biotech.ufl.edu (8.6.9/4.09)
id JAA05678; Mon, 29 Apr 1996 09:31:52 GMT




From: bozzola-at-siu.edu (John. J. Bozzola)
Date: Mon, 29 Apr 1996 15:16:50 -0600
Subject: Vacuum Pump Repair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130501adaae1a61df8-at-[131.230.97.69]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We have a 2 yr old Edwards 2-stage direct drive pump (Model E2M-12, serial
20372) that needs to be repaired (keeps blowing fuses). Anyone know of a
good repair place who might be able to handle this? Many thanks.


#############################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
#############################################################################






From: moninger :      moninger-at-emiris.iaf.uiowa.edu
Date: Mon, 29 Apr 1996 16:57:54 -0500 (CDT)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Mon, 29 Apr 1996, Joe Fu wrote:

} Greeting:
}
}
} Does anyone know the OSRAM Inc. telephone# & address? We need bulb
} replacement for our microscope.
} Yhank you.
}
} Joseph Fu
} NIST Rm.A117 Bldg.220
} Gaithersburg, Md. 20899
} Tel:301-975-3495
} e mail: jofu-at-enh.nist.gov
} Fax: 301-869-0822


Try:

ESV Inc.
(309)-347-6685

Tom

Thomas Moninger (moninger-at-emiris.iaf.uiowa.edu)
Central Microscopy Research Facility
85 EMRB University of Iowa
Iowa City, IA 52242 319-335-8142 FAX 319-335-8049





From: Beverly E Maleeff
Date: 29 Apr 96 11:41:55 EDT
Subject: May '96 PSM Meeting Notice

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9604291839.AA1097-at-pho903.sbphrd.com}
To: microscopy {microscopy-at-Sparc5.Microscopy.Com}
{Beverly_E_Maleeff-at-sbphrd.com}

Philadelphia Society for Microscopy Meeting Notice
May 1996


DATE: Thursday, May 9, 1996

PLACE: Laboratory for the Research of Science and Materials
(LRSM) Building, 33rd and Walnut Street, Philadelphia, PA.
Parking is available behind the LRSM Building after 5:00 PM.

TIME: 5:30 Social hour, hosted by our meeting sponsor.

6:30 Dinner

7:30 PM Speaker:

Electron Microscopy in the Study of Copolymers:
Blend Miscibility and Shear-Induced Defects

Dr. Karen I. Winey
Department of Materials Science and Engineering
University of Pennsylvania
Philadelphia, PA

Polymers containing two or more types of monomer units are copolymers.
The different monomer units can be distributed along the polymer chain
in a variety of sequences which produce random, alternating, or block
copolymers. Both the sequence distribution and the relative amounts of
each monomer can produce significant changes in morphological, chemical
and mechanical properties. Two examples will be given from our studies of
copolymers which illustrate the value of electron microscopy.
(1) We have developed a method which combines transmission electron
microscopy, image analysis and mass balance to construct quantitative ternary
phase diagrams. (2) Shear-induced defects, reminiscent of kink bands,
have been imaged using low voltage scanning electron microscopy.
The use of electron microscopy as a tool in these studies will be discussed.


DINNER:

COST: Members $12.00 Student members $6.00 Non-members $15.00

MENU: Barbequed chicken breast
Barbequed beef ribs
Hamburgers with all the trimmings
Red bliss potato salad
Pasta salad with fresh vegetables

Brownies
Fresh fruit

Reservations will be taken by Ms. Pat Overend at the University of
Pennsylvania, 215/898-8337. Deadline for reservations will be Friday,
May 3. If you have any questions regarding the meeting please feel free to
contact
Rollin Lakis at 215/898-2013 or lakis-at-sol1.lrsm.upenn.edu.
Cancellations must be received by Ms. Overend no later than 5:00 PM,
May 3, 1996.





From: jmkrupp-at-cats.ucsc.edu (Jon Krupp)
Date: Mon, 29 Apr 1996 16:06:51 -0700
Subject: For Sale or Trade

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604292306.QAA19547-at-cats.ucsc.edu}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

These items are offered for sale or trade, I need a wheel for tripod polishing:

1. Polaron Series 6000 Coating System .

2. Kevex thin window, horizontal detector for JEOL SEG stage, detector
only, no electronics.

3. Link Be window, horizontal detector for JEOL SEG stage, Link PP and
bias unit, Dapple X-mate and MicroPlus system.

E-mail for details, each one needs more explanation if you are interested.


Jonathan Krupp
Microscopy and Imaging Lab
University of California
Santa Cruz, CA 95064
(408) 459-2477
FAX (408) 429-0146
jmkrupp-at-cats.ucsc.edu






From: John F. Mansfield :      jfmjfm-at-umich.edu
Date: Sun, 28 Apr 1996 23:51:51 -0400
Subject: Technicial Positions Open

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: jfmjfm-at-srvr5.engin.umich.edu
Message-Id: {v03006601ada9ed196746-at-[141.213.21.13]}
Mime-Version: 1.0
Content-Type: text/enriched; charset="us-ascii"
Content-Transfer-Encoding: quoted-printable

Please pass this information to anyone who is interested.


{bold} {fontfamily} {param} Times {/param} {bigger} {bigger} {bigger} TECHNICAL
SUPPORT OPPORTUNITIES {/bigger}


The University of
Michigan {/bigger} {/bigger} {/fontfamily} {/bold} {fontfamily} {param} Times {/para=
m} {bigger}


{bold} {bigger}

Department of Materials Science and Engineering {/bigger} {/bold} =20


Due to recent and pending retirements, three technical support
positions are open. The department seeks to hire a technical support
team for the research and teaching needs of the department. The
positions that are to be filled are outlined as follows:


{bold} {bigger} Engineer II/III, T-96-1172 {/bigger} {/bold}

Responsibilities include:

Maintenance, service and repair of instruments and research equipment
including electronics, vacuum systems, computer-based systems,
mechanical testing equipment. This includes a number of scanning &
transmission electron microscopes, an Auger system and X-ray
photoelectron spectrometer. Assist with the maintenance and service of=20
X-ray diffraction systems, including a computer controlled Rigaku
system. Installation and networking of computers, interfacing computers
to research instrumentation, installation of data acquisition packages
(e.g. LabVIEW) and programming data acquisition modules for such
packages. Review and improve the safety program of the department.=20
Minimum requirement: B.S. in engineering or equivalent, plus three
years of relevant experience.


{bold} {bigger} Instrument Analyst I/II, T-96-1171 {/bigger} {/bold}

Responsibilities include:

Maintenance and service of two scanning electron microscopes (one scope
is a field emission instrument) and two=20

X-ray diffraction systems, including a computer controlled Rigaku
system. Operate and train students on proper use of above mentioned
instruments. Operate and maintain departmental multimedia facilities
(presentation computers, VCRs, TVs, TV cameras, CD-ROM drives, etc.).
Assist with review and improvement of departmental safety program.=20
Minimum requirement: B.S. in science, engineering, or equivalent,
plus three years of relevant experience.


{bold} {bigger} Computer Systems Specialist I/II,
T-96-1176 {/bigger} {/bold}

Responsibilities include:

Assist in the planning, operation and maintenance of the hardware and
software (network and stand alone) for departmental computer systems.=20
Provide technical advice to faculty, support staff and graduate
students. Coordinate the implementation of problem resolutions for
system hardware, software and network attachments and other operational
needs; arrange for hardware repair; maintain a schedule of system
back-ups and recovery procedures; load, test and implement upgrades to
existing and/or new applications; provide user training on upgraded or
new applications. Evaluation and acquisition of software packages.=20
Installation and networking of computers, interfacing computers to
research instrumentation, installation of data acquisition packages
(e.g. LabVIEW) and programming data acquisition modules for such
packages. Operate and maintain departmental multimedia facilities
(presentation computers, VCRs, TVs, TV cameras, CD-ROM drives, etc.).=20
Assist with review and improvement of departmental safety program.=20
Minimum requirement: B.S. in science or engineering, plus two years of
relevant experience.



Interested individuals should submit 2 copies of resume to: Employment
Services, The University of Michigan, G300 Wolverine Tower, 3003 S.
State Street, Ann Arbor, MI 48109-1281. Reference the Job # listed
next to the job title above. A non-discriminatory, affirmative action
employer.

{bigger}








{/bigger} {/bigger} {/fontfamily}

John Mansfield

North Campus Electron Microbeam Analysis Laboratory

417 SRB, University of Michigan

2455 Hayward, Ann Arbor MI 48109-2143 =20

Phone: (313)936-3352 FAX (313)936-3352

Email: jfmjfm-at-engin.umich.edu

URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html





From: slakmon-at-scottscientific.com (P. Slakmon)
Date: Mon, 29 Apr 1996 15:39:45 -0400 (EDT)
Subject: Conference & Workshop Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


S C O T T S C I E N T I F I C


On the behalf of The International Society of Molecular Morphology

Would like to take the opportunity to announce the

FOURTH INTERNATIONAL CONFERENCE & WORKSHOP
on
MOLECULAR MORPHOLOGY

June 3-4, 1996 - Conference
June 5 -6, 1996 - Workshop
in

Montreal, Canada

CONFERENCE FEATURING
Advances in Principles, Techniques and
Applications in Research and Diagnosis of:

- In Situ PCR
- In Situ Hybridization
- Immunohistochemistry
- Immunogold-Silver Staining
- Immunogold Electron Microscopy
- Microwave Immunohistochemistry
- Atomic Force Probe Microscopy
- Confocal Microscopy
- Antigen Retrieval
- Image Analysis

Call for Abstract Submission
DEADLINE: APRIL 15, 1996

The abstract should be typed single space on white paper within 9 x 7 inch
(23 x 18 cm) typed space. Total of two pages per abstract. Photographs and
references may be included. Please follow the style of CELL VISION, in
which the Proceedings will be published. Abstracts should be submitted in
duplicate.

THREE HANDS-ON WORKSHOPS (at the laboratories of Dept. of Anatomy,
University of Montreal)
1 In Situ PCR and In Situ Hybridization
2 Immunogold EM and Immunogold-Silver Staining
3 Microwave Fixation, Antigen Retrieval and Immunohistochemistry

June 2, 1996 (Sunday), 6.30-9.30 pm:
An optional preparation lecture on "Molecular Biology for the Uninitiated"

ORGANIZING COMMITTEE
CO-CHAIRMEN
Jiang Gu, M.D., Ph.D.
Deborah Research Institute
Browns Mills, New Jersey, USA

Moise Bendayan, Ph.D.
University of Montreal
Montreal, Canada

MEMBERS
Virginia Anderson, M.D.
Health Science Center at Brooklyn
State University of New York
Brooklyn, New York, USA

Gerhard Hacker, Ph.D.
Institute of Pathology
General Hospital, University of Salzburg
Salzburg, Austria

Lawrence DeBault, Ph.D.
Oklahoma University Health Center
Oklahoma City, Ok, USA

Shahla Masood, M.D.
University of Florida Health Science Center
Jackonsville, Florida, USA

Robert Day, Ph.D.
University of Montreal
Montreal, Canada

David Kersten, M.D.
Ealing
London, UK

----------------------------------------------------------------------------
--------
ADVANCED REGISTRATION FORM
(please print and use photocopies for additional forms)

NAME ________________________________

PHONE _________________ FAX __________________

ADDRESS __________________________________________

_________________________________________

CHOICE OF WORKSHOP (circle one): 1 2 3
----------------------------------------------------------------------------
-
ADVANCED REGISTRATION FREE (for one of three):
- For the two-day conference $250 (US)
- For the two-day workshop $350 (US)

Make check payable to CELL VISION. A 15 % discount for members of the
"International Society of Molecular Morphology" and students (with proven
ID). A 15% discount will be reimbursed upon becoming a member of the
society before or at the conference.
Please book the hotel room directly by calling The Best Western Hotel in
Montreal, Canada (800) 361-3000 or Fax (514) 861-4089. You can obtain a
special discounted room rate at $79 (Canadian) per day (rate includes
breakfast) by identifying yourself as a participant of the conference/workshop.
Student dormitory available at University of Montreal (15-minute subway
transporation to conference location, 5 minute walk to workshop location) at
$35 (Canadian) per day by calling (514) 343-6531.

Send abstract, registration form, and registration fee to:
CELL VISION-JOURNAL OF ANALYTICAL MORPHOLOGY,
EDITORIAL OFFICE, DEBORAH RESEARCH INSTITUTE
1 Trenton Road, Browns Mills
NJ 08015-1799, USA
Phone: (609) 735-0477
Fax: (609) 735-0478


For further information please direct your inquiries by email to:
morphology-at-scottscientific.com
_______________________________________________
SCOTT SCIENTIFIC
P.O. Box 66552 Station Cavendish,
Montreal, Quebec, H4W 3J6, Canada

Telephone: 514-485-2309 Fax: 514-485-9931
Voice Mail: 514-888-6509

WWW Site: http://www.ScottScientific.com

E-Mail: slakmon-at-scottscientific.com
info-at-scottscientific.com
sales-at-scottscientific.com
admin-at-scottscientific.com
_________________________________________________


_______________________________________________
SCOTT SCIENTIFIC
P.O. Box 66552 Station Cavendish
Montreal, Quebec, H4W 3J6, Canada

Telephone: 514-485-2309 Fax: 514-485-9931
Voice Mail: 514-888-6509

WWW Site: http://www.scottscientific.com

E-Mail: info-at-scottscientific.com
sales-at-scottscientific.com
admin-at-scottscientific.com
links-at-scottscientific.com
_______________________________________________





From: zaluzec-at-Sparc5.Microscopy.Com (Nestor J. Zaluzec)
Date: Mon, 29 Apr 1996 22:19:15 -0500
Subject: From the SysOp... Stick to Microscopy & Microanalysis!!!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604300315.WAA05275-at-Sparc5.Microscopy.Com}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

G'day Colleagues...

A Reminder: This forum should not to be used to post
non-Microscopy/Microanalysis
related announcements, requests, and/or informational items. There are
sufficient
number of public forums in which these issues may be freely and openly debated
and discussed to your heart's desire.

Please refrain from using the Microscopy ListServers for those purposes,
however, noble you may believe your motivations are....

There is no need for the ListServer community to echo this, as enough
bandwidth has
already been wasted, and enough mailboxes cluttered.

Nestor
Your Friendly Neighborhood SysOp






From: IAN HALLETT :      ihallett-at-MARCCRI.MARC.CRI.NZ
Date: Tue, 30 Apr 1996 17:21:52 GMT+1200
Subject: Observing living and dead cells: LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All

A colleague wishes to differentiate between living and dead cells.
Does anyone have a simple and reliable method (or if not simple at
least reliable).

The cells are from plant cell suspension cultures that have been
subjected to various environmental stresses.

Thanks in advance

Ian


Ian Hallett
HortResearch
Mt Albert Research Centre
Private Bag 92 169
Auckland, New Zealand
Fax 64-9-815 4201
Telephone 64-9-849 3660
EMail ihallett-at-hort.cri.nz




From: ZHENQUAN lIU :      zqliu-at-pccms.pku.edu.cn
Date: Tue, 30 Apr 1996 16:47:31 -0600 (CST)
Subject: E-mail Address wanted

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Nupop-Charset: English


Hi,

I would like to know the E-mail address of:

Professor E. A. Davis.

His mail address is:
Philosophical Magazine
Department of Physics and Astronomy
University of Leicester
University Road, Leicester LEI 7RH
England

If someone knows, please tell me.

Many thanks!

Zhen Quan Liu
---------------------------------------------------------------------
*Important!!!
ALL TELEPHONE NUMBERS in Beijing will be changed from May 8, 1996
from 7 digitals to 8 digitals!!!
The way of changing is by putting 6 infront of the initial number.
Therefore my tel since then will be
Tel: 6275 1427(office)
6275 3727(home)
Fax: 6275 1615
The codes for China and Beijing will not be changed, they still are:
China 86
Beijing 10
--------------------------------------------------------------------
** Sorry, my E-mail software cannot show me the address of
coming mails. Please tell me your E-mail address within
the text.
---------------------------------------------------------------------
Zhen Quan Liu (Ph.D) Tel:(86) 10 275 1427(Office)
Physics Building (86) 10 275 3727(Home)
EM Lab. Email: zqliu-at-pku.edu.cn
Physics Building Email(home) wl-at-ibmstone.pku.edu.cn
Peking University Fax (office): (86) 10 275 1615
Beijing 100871, China




From: David.Rothbard-at-ipst.edu (David Rothbard)
Date: Tue, 30 Apr 1996 08:07:25 -0500
Subject: Re: OSRAM Bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v021305a1adabc0833cb7-at-[199.77.235.102]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Joe:
Another source is Gray Supply Company. 800-238-2244. They have a wide range of projection of microscope bulbs and will send a catalogue.

--
David Rothbard
Institute of Paper Science and Technology






From: Fermin, Cesar :      fermin-at-TMC.Tulane.Edu
Date: 30 Apr 1996 07:06:40 -0500
Subject: Montage-got it-thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1381274340.4479-at-msmail.tmc.tulane.edu}

I already have the information needed and is enclosed below for anyone else
may need it.

The company was bought by the employees. Their new name is Montage
Graphics, the address is 2730 Scott Boulevard, Santa Clara CA 95050. The
phone is Sales - 800-416-4166, Email sales-at-montagegraphics.com, Technical
support 408-654-0684, Tech email tech-at-montagegraphics.com

Mei Lie Wong
Department of Biochemistry
HHMI-UCSF
Ph. 415-476-4441 Fax 415-476-1902
email wong-at-msg.ucsf.edu




From: rnessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Tue, 30 Apr 1996 09:21:52 -0500 (CDT)
Subject: Re: Vacuum Pump Repair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Mon, 29 Apr 1996, John. J. Bozzola wrote:

} We have a 2 yr old Edwards 2-stage direct drive pump (Model E2M-12, serial
} 20372) that needs to be repaired (keeps blowing fuses). Anyone know of a
} good repair place who might be able to handle this? Many thanks.
}
}
} #############################################################################
} John J. Bozzola, Ph.D., Director
} Center for Electron Microscopy
} Southern Illinois University
} Carbondale, IL 62901-4402
} U.S.A.
} Phone: 618-453-3730
} Fax: 618-453-2665
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/departments/shops/cem.html
} #############################################################################
}
We had a Balzers rotary pump rebuilt by Atlantic Vacuum Repair
-at-201-623-1115. They did a great job, and I would recommend that you give them a
call. They are at 96 Harper St., Newark, NJ. 07114. I'm one satisfied customer.


Randy Nessler
rnessler-at-emiris.iaf.uiowa.edu






From: Stefan Andreatta :      Stefan.Andreatta-at-uibk.ac.at
Date: Tue, 30 Apr 1996 16:29:02 +0100
Subject: Re: Observing living and dead cells: LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ian

I have not tried differentiating between living and dead eucaryotic
cells, but maybe it would be a good idea to contact Molecular Probes
Europe BV. They are selling a number of live/dead fluorescence
staining kits for both procaryotic and eucaryotic cells.

(unfortunately they are not going to pay me for posting their address)
Molecular Probes Europe BV, 2333 AA Leiden, The Netherlands
Phone: +31-71-233378; Fax: +31-71-233419

Stefan


* Stefan Andreatta
* Institute of Zoology and Limnology
* University of Innsbruck; Austria







From: Elinor Solit :      cambrex-at-world.std.com
Date: Tue, 30 Apr 1996 10:54:50 -0400 (EDT)
Subject: Re: for microscope manufacturers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


James,

Try the following companies and people:

Patrick Drexel or Tom Freda at Prior Scientific, 800-877-2234

Irv Toplin at Carl Zeiss, 800-356-1090

Jim Dutkiewctz at Meiji, 800-832-0060.

All these references feature products in The Microscope Book. The Spring
Edition is on press. Would you like a copy?

Please call us if there is anything more we can do to assist.

Elinor Solit, Director of Publications,
The Cambrex Group
800-440-0311

On Fri, 26 Apr 1996, James S MArtin wrote:

} Several months ago I posted a request for microscope manufacturers to
} contact me with information on costs for epi-fluorescence upgrades to
} polarizing light microscopes. The information will be presented at a
} national conference.
}
} So far, I have information only from Olympus. Anybody else like to be
} represented?
}
} James Martin
} Director of Analytical Services and Research
} Williamstown Art Conservation Center
} Williamstown, MA
}




From: Wil Bigelow :      Wil_Bigelow-at-mse.engin.umich.edu
Date: 30 Apr 1996 12:46:40 -0400
Subject: RE-VacPmpRepair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1381253843.32880-at-mse.engin.umich.edu}

Subject: Time: 7:36 AM
OFFICE MEMO RE:VacPmpRepair Date: 4/30/96

Most manufacturers of vacuum pumps will provide repair and maintenance
services for their pumps. Usually, to save time in getting you back into
operation, they will make an exchange with you, immediately shipping you a
used, but fully reconditioned pump, for the one you send in for repair.
Having the pump serviced by the manufacturer should be about the most
reliable way of ensuring that it receives proper attention.
The phone number of the Edwards service unit is 716-695-6354

The number of the service





From: hua-at-junction.ucsb.edu (Wu Xuehua)
Date: Tue, 30 Apr 1996 11:13:40 -0800
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604301814.LAA25533-at-surface.ucsb.edu}

unsubscribe please
Xuehua Wu
Materials Department
University of california, Santa Barbara
CA 93106
Tel: 805 893 8523





From: Tony King , INTERNET:tking-at-vacgen.fisons.co.uk
Date: Mon, Apr 29, 1996, 12:40 PM
Subject: Re: Peltier vs adabiatic CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I passed on the inquiry regarding Peltier vs adabiatic coolong of CPDs to Tony
King, Product Specialist at VG Microtech, who manufacture the Polaron range of
critical point dryers which we sell. This is his reply:


RE: Re: Peltier vs adabiatic CPD


One of the main reasons for changing from adibiatic cooling was due to
health and safety, 10% CO2 in air can cause permanent brain damage.
Secondly the cooling of a large vessel suitable for a number of
samples can be slow and difficult to control. If
you use a peltier cooled device the big advantage is that by reversing
the polarity the device acts as a heater with a limited range so it
is safe and efficient for this very application.

The only draw back that we have seen is that in a warm area the
excess heat must be removed by additional cooling. We used air
previously but again this was inefficient so we went to water.

As soon as you add peltier cooling and heating you can run a 3 term
controller and any one that wants to be clever can heat or cool
in precise stages!.

I hope this helps


Regards,

Tony King
Product specialist
VG Microtech/ Polaron range

Tel: +44 (0)1825 746251
Fax: +44 (0)1825 768343

Disclaimer:
The views and opinions expressed are not necessarily
those of Fisons plc or VG Microtech.







From: cxb41-at-po.CWRU.Edu (Christine H. Block)
Date: Tue, 30 Apr 1996 15:13:31 -0400
Subject: Point Source Bulbs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199604301913.PAA13964-at-roo.INS.CWRU.Edu}

I am looking for a supplier for a point source bulb for
a DURST ENLARGER... any suggestions?

Thanks


--
\\\
(o o)
*Limbic Lady* -------oOO--(_)--OOo-----




From: EmLab :      EMLAB-at-OPUS.MCO.EDU
Date: Tue, 30 Apr 1996 13:31:22 -0500 (EST)
Subject: Re: Vacuum Pump Repair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear John,
Try letting your maintance shop do it. I was informed by my old job maintance
men that a direct drive pump is good for about 2-3 rebuilds. The belt driven
pumps can take about 8-10 rebuilds.

Best of luck,
Ed Calomeni
Medical College of Ohio
Toledo, OH 43699
emlab-at-opus.moc.edu




From: Eric Steel :      steel-at-enh.nist.gov
Date: Tue, 30 Apr 1996 09:49:45 -0400 (EDT)
Subject: Re: Vacuum Pump Repair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Attached is a technical meeting announcement for anyone interested

MAMAS
(Mid-Atlantic Microbeam Analysis Society)
Meeting
at the
National Institute of Standards and Technology
Gaithersburg, MD
on
Thursday, May 9, 1996
10:30 am- 2:15 pm
Lecture Room D, Administration Bldg.

10:30 Coffee and Doughnuts

10:45 Joe Geller, Geller Microanalytical Laboratory, Standards
for the SEM Laboratory, Preparation for ISO-9000

12 noon Lunch

1:15 John Armstrong, NIST, Evaluating and Choosing Standards
for Quantitative Microbeam Analysis

For more information, contact Ryna Marinenko (301)975-3901,
FAX(301)216-1134, email:ryna.marinenko-at-nist.gov


Eric B. Steel, Leader e-mail: eric.steel-at-nist.gov
Microanalysis Research Group Office: 301-975-3902
N.I.S.T. FAX: 301-216-1134
Bldg. 222/Rm A113
Gaithersburg MD 20899





From: cmurphy-at-GGPL.ARSUSDA.GOV
Date: Tue, 30 Apr 1996 16:15:22 -0400
Subject: Many Thanks

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would just like to thank all ( and there were many ) scientist that took the time to send me information regarding the frozen 4%
osmium. The information came so quickly that I was able to carry out an experiment with complete confidence of the outcome. By
the way it was unanimous. The OSO4 is O.K. if it has been frozen. Thanks again. Sincerely Charlie Murphy




From: rutledge phil :      prutle1-at-gl.umbc.edu
Date: Tue, 30 Apr 1996 10:32:25 -0400 (EDT)
Subject: Re: your mail

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Mon, 29 Apr 1996, Joe Fu wrote:

} Greeting:
}
}
} Does anyone know the OSRAM Inc. telephone# & address? We need bulb
} replacement for our microscope.
} Yhank you.
}
} Joseph Fu
} NIST Rm.A117 Bldg.220
} Gaithersburg, Md. 20899
} Tel:301-975-3495
} e mail: jofu-at-enh.nist.gov
} Fax: 301-869-0822
}
Joe:
Have you tried Bulbman Inc. in Reno, Nevada? They have had every
type of bulb I've needed. The cost is usually a fraction of what the
manufacturer charges. So far they have been the least expensive of any
other people I've tried. The people are pretty good handling customers
and their needs. Excellant to deal with.
Their number is: 1-800-648-1163
Fax: 1-800-548-6216
One time I had to replace a bulb, the bulb had no recognizable numbers
to give them. I faxed them the type and model of the microscope and they
were able to give me the right bulb.
Hope this helps.

Peace,

Phil

p.s. They have a pretty quick turn around time.




From: John M. Libert :      jlibert-at-cpcug.org
Date: Tue, 30 Apr 1996 17:19:58 -0700
Subject: Re:

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {3186AE2E.1F0D-at-cpcug.org}

Try OPELCO, Inc.
703-471-0080 ext. 544


Joe Fu wrote:
}
} Greeting:
}
} Does anyone know the OSRAM Inc. telephone# & address? We need bulb
} replacement for our microscope.
} Yhank you.
}
} Joseph Fu
} NIST Rm.A117 Bldg.220
} Gaithersburg, Md. 20899
} Tel:301-975-3495
} e mail: jofu-at-enh.nist.gov
} Fax: 301-869-0822




From: Bob Citron :      Bob_Citron-at-cc.chiron.com
Date: 4/30/96 2:42 AM
Subject: Observing living and dead cells: LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear All

A colleague wishes to differentiate between living and dead cells.
Does anyone have a simple and reliable method (or if not simple at
least reliable).

The cells are from plant cell suspension cultures that have been
subjected to various environmental stresses.

Thanks in advance

Ian


Ian Hallett
HortResearch
Mt Albert Research Centre
Private Bag 92 169
Auckland, New Zealand
Fax 64-9-815 4201
Telephone 64-9-849 3660
EMail ihallett-at-hort.cri.nz





From: John F. Mansfield :      jfmjfm-at-umich.edu
Date: Tue, 30 Apr 1996 00:20:08 -0400
Subject: WWW Topical Conference in Minneapolis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: jfmjfm-at-srvr5.engin.umich.edu
Message-Id: {v03006606adab409f3583-at-[141.213.21.13]}
Mime-Version: 1.0
Content-Type: text/enriched; charset="us-ascii"

Here is a reminder and some clarifying information.


The Microbeam Analysis Society is sponsoring their first Topical
Conference immediately prior to the Microscopy and Microanalysis 96
meeting in Minneapolis this year. The Topical Conference is entitled:


{bold} "Microscopy & Microanalysis Resources on The World Wide Web"

(http://www.microanalysis.org/mas/topicalconf96.html

http://www.microanalysis.org/mas/regtopconf.html)

{/bold}

It will be held in The Minneapolis Convention Center on


SATURDAY THE 10TH of AUGUST 1996.


Note that the ad in Microscopy Today contained a typo stating that the
meeting was to be on the Sunday. The meeting has been placed on
SATURDAY to avoid any conflict with the M&M Tutorials.


The starting time will be 9:00am and the conference will last all day.


The morning will be a series of presentations which will cover the
following:


1. Basic introduction to the Web

2. What is available on the Web generally.

3. What is available on the Web for Microscopy & Microanalysis

4. How to create your own Web Page.


Invited Presenters include:

Greg Meeker - U.S. Geological Survey.

Marc De Graef - Carnegie Mellon University.

Darcy Clark - University of Michigan (formerly University of
Queensland).

John Mansfield - University of Michigan.


The afternoon will be focussed on a hands-on workshop where there will
be a minimum of 15 computer workstations connected to the Internet (via
a T1 line)

for attendees to use. The morning's presenters will be available to
advise attendees on all aspects of accessing and using the Web.


Attendance is FREE to MAS members

Attendance is $35 for Non Members

If you join The Microbeam Analysis Society you may attend free,
membership is $25 per year.


You may register by:

1. Filling out the card that was contained in the last issue of
Microscopy Today

and mailing it in to the address noted on the card.

2. You may register electronically if you are a member. Connect to:

http://www.microanalysis.org/mas/regtopconf.html

3. Non members may complete the form at:

http://www.microanalysis.org/mas/regtopconf.html, print it and mail it
to the address supplied in the form.

4. Members & Nonmembers may send their:

Name

Full address

Business phone

FAX

email address

and an indication of their computer and Web expertise to John
Mansfield

at the address at the end of this message.

Non-Members should enclose a check for $35 (drawn on a US bank) made
out to The Microbeam Analysis Society.



For future updates keep monitoring:

http://www.microanalysis.org/mas/topicalconf96.html

or send mail to John Mansfield (jfmjfm-at-umich.edu)


Thanks.




John Mansfield

North Campus Electron Microbeam Analysis Laboratory

417 SRB, University of Michigan

2455 Hayward, Ann Arbor MI 48109-2143

Phone: (313)936-3352 FAX (313)936-3352

Email: jfmjfm-at-engin.umich.edu

URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html





From: kingsley-at-zephyr.nrlssc.navy.mil (KingsleyMcCrocklin)
Date: 4/30/96
Subject: unsuscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This message was sent by Chameleon
-------------------------------------





From: kingsley-at-zephyr.nrlssc.navy.mil (KingsleyMcCrocklin)
Date: 4/30/96
Subject: unsuscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This message was sent by Chameleon
-------------------------------------





From: tphillips-at-biosci.mbp.missouri.edu (Tom Phillips)
Date: Tue, 30 Apr 1996 13:01:08 -0500
Subject: B&W photos of Cy5

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v01540b00adac0552a116-at-[128.206.15.189]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I have been using Cy5 with great luck on my confocal but now need to take
some B&W photos using my epi-fluorescence rig. I know something like Tri-X
pan would be virtually insensitive to the far red emission of Cy5. Does
anybody have practical experience with Tmax and Cy5 or is there a better
choice of B&W film for this purpose. Thanks.


Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(314)-882-4712 (voice)
(314)-882-0123 (fax)






From: !Microscopy-request-at-Sparc5.Microscopy.Com (Robert Kayton,MAC,CROET)
Date: Tue Apr 30 10:54:50 -0400 1996
Subject: Re: for microscope manufacturers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sender: kayton-at-ohsu.edu (Robert Kayton)
Message-Version: 2
} To: Microscopy-at-Sparc5.Microscopy.Com

James,

Try the following companies and people:

Patrick Drexel or Tom Freda at Prior Scientific, 800-877-2234

Irv Toplin at Carl Zeiss, 800-356-1090

Jim Dutkiewctz at Meiji, 800-832-0060.

All these references feature products in The Microscope Book. The Spring
Edition is on press. Would you like a copy?

Please call us if there is anything more we can do to assist.

Elinor Solit, Director of Publications,
The Cambrex Group
800-440-0311

On Fri, 26 Apr 1996, James S MArtin wrote:

} Several months ago I posted a request for microscope manufacturers to
} contact me with information on costs for epi-fluorescence upgrades to
} polarizing light microscopes. The information will be presented at a
} national conference.
}
} So far, I have information only from Olympus. Anybody else like to be
} represented?
}
} James Martin
} Director of Analytical Services and Research
} Williamstown Art Conservation Center
} Williamstown, MA
}




From: Ian J. Davies :      davies-at-asuka.nal.go.jp
Date: Thu, 1 May 1986 09:48:05 +0000
Subject: OM - Problem with Olympus SZH

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {9605010047.AA03572-at-asuka.nal.go.jp}
Comments: Authenticated sender is {davies-at-asuka.nal.go.jp}

I work at a research institute in Japan in the field of materials
science. I have experience in TEM and SEM but recently have been
trying to take micrographs using an Olympus SZH optical microscope.
This microscope has a "zoom" lens from nominally 7.5 to 64.

My problem is that this microscope appears to exhibit something
similar to astigmatism even at the lowest magnification. We also have
an Olympus STM5-BD which works perfectly. With the SZH I have tried
to realign the optical system but there is little improvement.
Another department in my institute also has an SZH with exactly the
same problem.

Does anyone know if this problem is common with the SZH
microscope and how it may be corrected.

Please could you send any thoughts and suggestions to me at
DAVIES-at-ASUKA.NAL.GO.JP

Any ideas would be gratefully received. Thank you.

Dr. Ian Davies




From: William Tivol :      tivol-at-wadsworth.org
Date: Tue, 30 Apr 1996 17:16:05 -0400 (EDT)
Subject: Re: Peltier vs adabiatic CPD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} If
} you use a peltier cooled device the big advantage is that by reversing
} the polarity the device acts as a heater with a limited range so it
} is safe and efficient for this very application.
}
} The only draw back that we have seen is that in a warm area the
} excess heat must be removed by additional cooling. We used air
} previously but again this was inefficient so we went to water.
}
} As soon as you add peltier cooling and heating you can run a 3 term
} controller and any one that wants to be clever can heat or cool
} in precise stages!.
}
Dear Tony,
We have a darkroom rinsing tank which is heated/cooled by peltier
devices. One caution with using these devices is that most of the commer-
cial controllers are duty-cycle type, i.e., they turn from full off to full
on for a fraction of the time which depends on the difference between the
set point and the temperature being controlled. This kind of controller is
harmful to peltiers. Our electronics shop designed a controller which puts
out a voltage which is proportional to the set-point-temperature difference,
so it adjusts slowly as the temperature reaches the set point and never has
sudden changes in output. With this kind of controller, the peltiers have
lasted many years without apparent damage.
Yours,
Bill Tivol




From: saz1-at-ix.netcom.com (Steven Zenk )
Date: Tue, 30 Apr 1996 21:10:04 -0700
Subject: Subscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please enter my subscription to your service. Please use the following
e-mail address: szenk-at-kevex.com. Should you have any questions,
please contact me.

Thanks for your continued efforts to satisfy the needs of the
microscopy consortium.




MicroscopyListserver Archive Email Extraction Software Version NJZ07060908