From: battista.calvieri-at-mailbox66.utcc.utoronto.ca (Battista Calvieri)
Date: Sun, 2 Mar 1997 21:32:42 -0500
Subject: Conversion the AN10000

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The software for a conversion of AN10000 to Windows has already been written
by a Toronto based company called Mektech. They also make a hardware to
interface the AN10000 pulse processor to a PC. You can find them at
mektech-at-visionol.net
The Microscopy Imaging Laboratory at the University of Toronto Dept.
of Medicine. Has purchased the interface for the AN10000 to PC and the
software from Mektech. The software is easy to operate and we have
found it to be very reliable. Mektech has allowed us to upgrade our
system at a fraction of the cost of that of a new system.


Battista Calvieri
Manager of Microscopy Imaging Labarotory
University of Toronto










From: Richard Beanland +44 1327 356363 :      richard.beanland-at-gecm.com
Date: Mon, 03 Mar 1997 10:06:28 +0000 (GMT)
Subject: Re: "Sprinkled" EM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Not even the suggested sophistication of a "halon" fire protection
system will render your EMs immune from the dreaded "water shower"
experience. Overnight flooding of the lab floor above us caused water
to find its way to (and then through) the smoke/ thermal detectors that
we had smugly positioned directly over the instrument for our halon gas
quench system, cascading water onto a powered TEM/STEM/EDS facility. By
the time I was called in, the TEM was lit up like a Xmas tree,
accompanied by the delicate aroma only smoldering electronics can
generate.

Unfortunately, similar water showers in the adjacent SEM room had missed
the console, and the instrument was too heavy to push under the water
stream (we were trying to get a new SEM at the time!).

Note: Our fire detectors have since been relocated, and the "halon" has
been replaced (by Govt decree) with environmentally more friendly (less
unfriendly?) FM200.

Eric Bradley
BHP Steel
Port Kembla NSW Australia
bradley.eric.eg-at-bhp.com.au


---------------------------------------

Some months ago a gate valve went in the cooling pipes for our building, which
created a spectacular waterfall in the corridor. Since an insurance claim is
the easiest way of getting new equipment here, we were quite hopeful, but like
Eric Bradley, we found that three strong people were unable to push either our
17 year old TEM, 25 year old Auger kit, or even the little 18 year old SEM
under the water.




Richard Beanland

GEC-Marconi Materials Technology,
Caswell,
Towcester,
Northants
NN12 8EQ





From: morita3-at-chaos.bio.sci.osaka-u.ac.jp (=?ISO-2022-JP?B?GyRCPzlFRD4tO0sbKEI=?=)
Date: Mon, 3 Mar 1997 18:46:28 +0900
Subject: Help : neuron number count by fluorescrnce probe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all

I have a problem about cell number count on agrregated neurons with
an enzyme treatment.
I want to check a viability after the enzyme treatment. It is usually
available to count a cell
number under phase contrast microscopy, but light was so reflected
that each neuron could not be
identified. I would, therfore, try to apply a fluorescence dye,for
example, Hoechst33342,
that bind to a minor grove of DNA and is heared to be applicable to a
native cell. But I wonder
it will work well. I would like to get more detailed information
about image analyze of cell
number under a microscopy.
Thanks a lot.


Masahito Morita
Laboratory of Compararive physiology
Faculty of Science
Osaka University











From: Damian_Neuberger_at_RLT013-at-ccmailgw.mcgawpark.baxter.com
Date: Mon, 3 Mar 1997 08:42:54 -0600
Subject: re: gloves

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Richard Easingwood wrote:

} I am trying to compile a list of what type of disposable gloves
should be } used with what chemicals found in our EM unit.
[snip]
}
} Organic solvents:
}
} acetone: butyl? or latex
} ethanol: latex
} methanol: latex
} chloroform: (PVA if avail) or nitrile (double glove) }
[snip]

Dear Richard,
I'd be very surprized if latex gloves would hold up to acetone;
when I've exposed them, they get tacky. Polyethylene gloves should
hold up to the organic solvents you listed (and many others).
Besides, they smell better than latex gloves.
Yours,
Bill Tivol

Richard, and others interested;

We use 100% Nitrile gloves as they stand up to all the chemicals we
use in the EM lab, We use the product available from Best
Manufacturing Company, Menlo, GA 800-241-0323. I'm sure there are
others just as suitable.

Disclaimer: We have no financial or other interest in the company.

Damian Neuberger




From: Nestor J. Zaluzec :      zaluzec-at-Sparc5.Microscopy.Com
Date: Mon, 3 Mar 1997 08:58:09 -0500
Subject: MSA/MAS spectrum file format

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sorry Colleagues,

But I've been on the road for awhile and hence
did not have time to catch up on things. Here's the info
that a few people have asked for as I did not see anyone
else post an answer.

The MSA/MAS Spectrum File Format Information & Examples
can be downloaded from the following FTP Site.


Host: ftp.amc.anl.gov
UserID: anonymous
Password: Your Email Address

Follow the Directory tree as follows:

**} ANLSoftwareLibrary
*****} 2-EMMPDL
********} Xeds
**********} EMMFF

Everything is in the EMMFF ( "E"MSA MAS File Format)
directory, including examples and documentation.

Version 1.0 is the current revision level. We (The MSA/MAS
Standards Committee) have not had a need to update it
beyond that point.

I have not gotten around to creating a WWW page for this.
If there is enough interest I will do it. However if you have
access to the WWW you can get it via FTP. It is also
routinely available in the Computer Workshop at the
annual Microscopy & Microanalysis Meeting .
( see http://www.msa.microscopy.com for this
years meeting information )

Cheers... Nestor
Your Friendly Neighborhood SysOp


Here is the Abstract File for that directory.

--------
Title :EMMFF V. 1.0
Keywords :XEDS,EELS,AES,WDS,CLS,GAM,XRF,PES
Computer :IBM, MAC, DEC
Operating System :ALL
Programming Language :Fortran 77
Hardware Requirements :None
Author(s) :EMSA/MAS TASK FORCE
Ray Egerton ,Charles E. Fiori ,John A. Hunt,
Mike S. Isaacson,Earl J. Kirkland ,Nestor J. Zaluzec
Correspondence Address :R.F. EGERTON-CHAIRMAN
University of Alberta
Dept. of Physics
Edmonton, Alberta, Canada, T6G2J1
Abstract:

A simple format for the exchange of digital spectral data is
presented, and proposed as an EMSA/MAS standard. This format is readable by
both humans and computers and is suitable for transmission through various
electronic networks (BITNET, ARPANET), the phone system (with modems) or on
physical computer storage devices (such as floppy disks). The format is not
tied to any one computer, programming language or computer operating system.
The adoption of a standard format would enable different laboratories to
freely
exchange spectral data, and would help to standarize data
analysis software. If equipment manufacturers were to support a common format,
the microscopy and microanalysis community would avoid duplicated effort in
writing data-analysis software. This version of EMSAMASFF contains two
subroutines which read and write spectral data files Version 1.0 data format.
The data are stored as simple ASCII characters at a user defined number of
columns per line for the length of the data file. The spectral data is
preceeded by a series of header lines, which tell the user about the
parameters of the spectrum. The header lines are identified by the first
character in the line being the symbol (#) followed by a descriptor and if
appropriate its units. An example of a data file format can be found in the
EMSAMASFF.DOC file.
------------------------------------------------------------------------------








From: ebs-at-ebsciences.com
Date: Mon, 03 Mar 1997 10:03:51 EST
Subject: "wehnelt"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear fellow microscopists,

You never know what you will find in your e-mail "in" basket. If any of you
know the derivation of the word "Wehnelt", please respond to this gentleman
directly at:

gustavo waehneldt {guswae-at-tricom.net}
} Subject: wehnelt

} The subscriber is obviously a relative of this "wehnelt" you mention. I
} belong to the Brazilian side of the family, hence the different graphy of
} the name. Can you be kind enough to give more info on this man? I knoe there
} was an Arthur Waehneldt who dedicated himself to physics. Are we
} talking about the same person? Pls. send all the details you may have. I`m
} trying to trace member oof the family ald your cooperation will be very
} wellcome. Thanking you in advance, I remain.
} Gustavo
} gustavo waehneldt
} guswae-at-tricom.net
********************************
Energy Beam Sciences, Inc.
Adding Brilliance To Your Vision
ebs-at-ebsciences.com
http://www.ebsciences.com/
********************************





From: beebed-at-ere.umontreal.ca (Dwight Beebe)
Date: Mon, 3 Mar 1997 10:10:32 -0400
Subject: Thanks for Ultracut assistance

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Good morning,
I would like to thank everyone who so helpfully provided
information on servicing the stereo pod on my Ultracut:

Lou Ann Miller, Julian Smith III, Normand Laurier, Allan Mitchell,
Kevin Hlacrow, Sara Miller, Dan Focht, Geoff McAuliffe, and
Alexander Greene.

I appreciate, as always, having access to such a generous group of
people, who find the time to provide useful information. Thanks again!


Dwight Beebe E-mail: beebed-at-ere.umontreal.ca
Institut de recherche en biologie vegetale Voice: 514-872-4563
Universite de Montreal FAX: 514-872-9406
4101, rue Sherbrooke est
Montreal, Quebec H1X 2B2
Canada






From: Keith Ryan :      KPR-at-wpo.nerc.ac.uk
Date: Mon, 03 Mar 1997 15:44:33 +0000
Subject: Re: "Sprinkled" EM's -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

So it takes more than "three strong people" to push them under?
I'll bear that in mind!

We are still running an Edwards 12E6 coating unit coming up to 32 years
old (can anybody beat that?), it goes like a Trojan (great for ploughing,
farming joke!).

Regards - Keith Ryan
Plymouth Marine Laboratory, UK




From: Woody.N.White%650 :      Woody.N.White-at-mcdermott.com
Date: Mon, 3 Mar 1997 8:01:00 -0600
Subject: Re[2]: Disposable gloves

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Perhaps I state the obvious, but beware of (powder) lubricated gloves.
When I started SEM work (long, long ago), only powdered gloves were
used for specimen prep. - what a mess! It seemed to always be
"snowing" on samples sent to me from our failure analysis lab!
Although somewhat difficult to locate at the time, I managed to
convince all to go with powder-free. No more problem. Powder free
are now easy to find.....

Woody




From: Bob_Citron-at-cc.chiron.com (Bob Citron)
Date: 2/28/97 11:40 AM
Subject: "Sprinkled" EM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Peggy;

I know you have received alot of replies about this subject, but I can't resist
adding my experiences. I compare having a water sprinkler system in an EM lab to
gambling in Vegas; sometimes you win, sometimes you lose. We have a sprinkler
system directly over our Cambridge 200 SEM, and fortunately we have not had to
use it in 10 years - doesn't mean it won't happen tomorrow, though. We DID,
however, have our ceiling cave in on our scope during a heavy rain. The soggy
composite tiles were like chewed up spitwads all over the unit, and a steady
river of water was running from the roof directly into the keyboard and console.
Figuring that it was probably dead, I called the manufacturer for advise. We
fixed the leak, then opened the SEM up and dried everything the best we could
with blasts from a nitrogen tank. We allowed it to dry for 3 more days and
fired it up. No problems other than some sticky keys in the keyboard - pretty
amazing.

(BTW, this is not an advertisement for Cambridge/LEO. It's a real experience)

Regards,

-Bob
********************************
Bob Citron
Chiron Vision
555 W. Arrow Hwy
Claremont, CA 91711
ph: (909)399-1311
email: Bob_Citron-at-cc.chiron.com
********************************
______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

We recently moved our TEM and SEM from a building built in 1930 to one
built in 1995. In other words, to a building with sprinklers. Now each
microscope has a sprinkler directly overhead and I've been wondering if I
should cover the scopes at night -- just in case. Do other people have
this concern/problem? I haven't read anything about sprinklers
malfunctioning and going off but it is a bit unnerving seeing a shower head
directly over the scopes! I haven't seen covers in any of the catalogs
so assume a tarp is the way to go....Or is it?

Thanks for any feedback....

Peggy

Peggy Brannigan
Electron Microscopy
Floral and Nursery Plants Research Unit
National Arboretum

Bldg. 010A R.238
10300 Baltimore Avenue
Beltsville, MD USA20705

Phone: (301) 504-6097
Fax : (301) 504-5096
Email: brannign-at-asrr.arsusda. gov







From: Peter Jordan :      emsi-at-pe.net
Date: Mon, 03 Mar 1997 09:39:42 -0800
Subject: Need used SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {331B0CDE.34E7-at-pe.net}

Looking for used SEM, must have low Voltage (1KV or less) capability.




From: Carl Zeiss Microscopy :      micro-at-zeiss.de
Date: Mon, 03 Mar 1997 18:48:11 +0100
Subject: New high performance confocal Laser Scanning Microscope from Carl Zeiss

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

This world wide E-mail provides important information for everyone
involved in biomedical research.
We hope no one feels bothered by this mail. Spamming is not indented.

Dear researcher,

Carl Zeiss is very pleased to announce it's new high performance
confocal

Laser Scanning Microscope LSM 510,

designed for your best results in biomedical research.

(L)et's (S)ee (M)ore for applications, technical data and world wide
expert contacts

at http://www.zeiss.de/mi/limi_e/p4/lsm510_entry.htm

Sincerely Yours,

Carl Zeiss
Microscope Division (German headquarters)
D-07740 Jena, Germany
E-mail: micro-at-zeiss.de
Hotline: ++49 3641 64 1616
Fax: ++49 3641 63 3144
URL: www.zeiss.de





From: Carl Zeiss Microscopy :      mikro-at-zeiss.de
Date: Mon, 03 Mar 1997 18:48:11 +0100
Subject: New high performance confocal Laser Scanning Microscope from Carl Zeiss

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

This world wide E-mail provides important information for everyone
involved in biomedical research.
We hope no one feels bothered by this mail. Spamming is not indented.

Dear researcher,

Carl Zeiss is very pleased to announce it's new high performance
confocal

Laser Scanning Microscope LSM 510,

designed for your best results in biomedical research.

(L)et's (S)ee (M)ore for applications, technical data and world wide
expert contacts

at http://www.zeiss.de/mi/limi_e/p4/lsm510_entry.htm

Sincerely Yours,

Carl Zeiss
Microscope Division (German headquarters)
D-07740 Jena, Germany
E-mail: micro-at-zeiss.de
Hotline: ++49 3641 64 1616
Fax: ++49 3641 63 3144
URL: www.zeiss.de





From: Michael Shaffer :      mshaf-at-OREGON.UOREGON.EDU
Date: Mon, 03 Mar 1997 11:07:03 -0800
Subject: strontium in apatite and carbonate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


As an anthropological electron probe microanalytical (EPMA) project we are analyzing ancient and diagenetic bone for strontium, barium and other trace elements. We are seeing an unexpected (so what else is new) correlation of Sr with phosphate and the i
nverse with apatite replaced by carbonate. We are trying to eliminate analytical errors, and would like to throw this problem at the EPMA community for feedback ... my standard apatite with no Sr measures at detection, and we have double-checked for backg
round interferences, but only for elements suspected.
We are suspicious of not analyzing for a element which might be causing this problem ... TIA

cheerios, shaf

{\/} /\ {\/} /\ {\/} /\ {\/} cogito, ergo zZOooOM {\/} /\ {\/} /\ {\/} /\ {\/}
Michael Shaffer, R.A. - University of Oregon Electron Probe Facility
mshaf-at-oregon.uoregon.edu -or- mshaf-at-darkwing.uoregon.edu
http://darkwing.uoregon.edu/~mshaf/epmahome/




From: heiman-at-gmtme.com (Bob Heiman)
Date: 3/3/97
Subject: P+ Silicon Stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Does anyone know of a good preferential etch for P+
silicon? I've been using a 20:1 HNO3:HF stain. This stain
seems to delineate the N+ areas very well, but the P+
areas are not easily seen.

I would like to decorate the P+ with as little attack to
the N+ areas as possible.

Any help is greatly appreciated.

Regards,
Bob Heiman
-------------------------------------
E-mail: heiman-at-gmtme.com
Voice: 610-666-7950 x2855, FA Lab x2533






From: Simon Watkins :      swatkins-at-pitt.edu
Date: Mon, 03 Mar 97 14:16:02 -0500
Subject: drown the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199703031915.OAA14276-at-post-ofc04.srv.cis.pitt.edu}
To: microscopy {microscopy-at-Sparc5.Microscopy.Com}

-- [ From: Simon Watkins * EMC.Ver #2.5.02 ] --


As another tale of woe to follow up on Bob Citrons comment on soggy ceiling
tiles forming spit wads all over his SEM.

A few years back while I was working in Boston we tried to drown a 100CX by
backing up several gallons of water down an air handling chute directly on
top of the column (don't ask why or how, but it was one of the side effects
of being in the basement). Being on the ball, the operator recognised the
problem due to the high moisture content in the air and the splishy splashy
sound of water falling from the ceiling. He hit the magic red button to
shut the instrument down, and ran for help shouting as he ran "the
microscopes drowning, the micrscopes drowning".

As was the case with Bob we called the manufacturer, in this case the ever
helpful folks at JEOL, who agreed this was something of a problem but had no
immediate solution beyond "errrr, huh, interesting", so we dried off what we
could and left the thing off for a week. We turned it on and away it went,
as if nothing had happened. The only long term effect was some rusting of
the Allen bolts used to hold the column together, you would think that they
would be painted or stainless with all these "environmental" problems of
late!

my 2 cents

Simon


--
-------------------------------------------------
Simon C. Watkins Ph.D
Associate Professor
Director Center for Biologic Imaging
University of Pittsburgh
Pittsburgh PA 15261
tel 412-648-3051
fax 412-648-2004
-----------------------------------------------




From: Crossman, Harold :      crossman-at-OSI.SYLVANIA.com
Date: Mon, 3 Mar 1997 15:01:27 -0500
Subject: RE: P+ Silicon Stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The book, Handbook of Metal Etchants, has about a hundred pages of
etchant recipes for Si wafers, crystals. Good discussions on practices/
applications as well. Includes references.

Handbook of Metal Etchants
Walker, Perrin & Tarn, William H.
1991 CRC Press, Boca Raton, FL 33531
ISBN 0-8493-3623-6


Happy mixing!
------------------------------------------------
Opinions or statements expressed herein, rational or otherwise, do not
necessarily reflect those of my employer.

Harold J. Crossman
OSRAM SYLVANIA INC.
Lighting Research Center
71 Cherry Hill Dr.
Beverly, MA 01915
Phone: (508) 750-1717
E-mail: crossman-at-osi.sylvania.com

Our web sites: www.sylvania.com
www.siemens.com
--

"Crossman, Harold" {crossman-at-osi.SYLVANIA.com}




From: Peggy Brannigan :      brannign-at-asrr.arsusda.gov
Date: Mon, 3 Mar 1997 15:02:51 -0500
Subject: . Thank you, "wet" EMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Thanks to all who answered my recent question about covering EM's against
sprinkler damage! A lot of good points were made, including:

1. The threat is real (some real horror stories out there!)
2. It's not just sprinklers that threaten, it's floods from burst
plumbing, storms etc...
3. Tarps are inconvenient and possibly hazardous,
4. Curved plexiglass or fiberglass structures attached to the ceilings
over the microscopes were widely recommended (like salad bars)
5. Many suggested dismantling or plugging the sprinkler heads while
6. others were concerned about fire safety
7. A surprising number of drenched scopes fully recovered
8 which you should keep in mind if you actually push your scope under
the downpour in hopes of obtaining a new one.
9. but if you try anyway, it takes more than three people.

Thanks again!

Peggy

Peggy Brannigan
Electron Microscopy
Floral and Nursery Plants Research Unit
National Arboretum

Bldg. 010A R.238
10300 Baltimore Avenue
Beltsville, MD USA20705

Phone: (301) 504-6097
Fax : (301) 504-5096
Email: brannign-at-asrr.arsusda. gov







From: Michael OKeefe :      Michael_OKeefe-at-macmail.lbl.gov
Date: 3 Mar 1997 13:32:29 -0700
Subject: Why am I not surprised?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Subject: Time:1:27 PM
OFFICE MEMO Why am I not surprised? Date:3/3/97

Last week I received a message from Microscopy about "Little Jessica Mydeck".
A check at http://www.cancer.org/acs.html revealed:

Fraudulent Chain Letter
(This statement may be copied or reprinted by online users)

The American Cancer Society is greatly disturbed by reports of a
fraudulent chain letter circulating on the internet which lists the
American Cancer Society as a "corporate sponsor" but which has in
no way been endorsed by the American Cancer Society.

This letter appears to have started on America Online but has now
spread well beyond the online service. There are several variations
of this letter in circulation. The text of the original message reads
as follows:

LITTLE JESSICA MYDEK IS SEVEN YEARS OLD AND IS SUFFERING FROM
AN ACUTE AND VERY RARE CASE OF CEREBRAL CARCINOMA. THIS
CONDITION CAUSES SEVERE MALIGNANT BRAIN TUMORS AND IS A
TERMINAL ILLNESS. THE DOCTORS HAVE GIVEN HER SIX MONTHS
TO LIVE.

AS PART OF HER DYING WISH, SHE WANTED TO START A CHAIN
LETTER TO INFORM PEOPLE OF THIS CONDITION AND TO SEND PEOPLE
THE MESSAGE TO LIVE LIFE TO THE FULLEST AND ENJOY EVERY
MOMENT, A CHANCE THAT SHE WILL NEVER HAVE. FURTHERMORE,
THE AMERICAN CANCER SOCIETY AND SEVERAL CORPORATE SPONSORS
HAVE AGREED TO DONATE THREE CENTS TOWARD CONTINUING CANCER
RESEARCH FOR EVERY NEW PERSON THAT GETS FORWARDED THIS
MESSAGE. PLEASE GIVE JESSICA AND ALL CANCER VICTIMS A CHANCE.

IF THERE ARE ANY QUESTIONS, SEND THEM TO THE AMERICAN CANCER
SOCIETY AT ACS-at-AOL.COM

As far as the American Cancer Society can determine, the story
of Jessica Mydek is completely unsubstantiated. No fundraising
efforts are being made by the American Cancer Society in her
name or by the use of chain letters. Furthermore, the email address
ACS-at-AOL.COM is inactive. Any messages to the American Cancer
Society should be instead sent through the American Cancer
Society website at http://www.cancer.org.


Michael A. O'Keefe





From: fransisco black :      fransisco-at-hotmail.com
Date: Mon, 3 Mar 1997 16:20:46 -0800 (PST)
Subject: EDS related question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists:

I would like to get an explanation for the following observation
of quantitative analysis by EDS.

I am looking at a Si wafer with 100nm of SiO2 layer on top.
from a SEM. Electron beam (~8 kev) hit the surface along the
surface normal (0 deg tilt). As I increase the sample tilt
(upto 45 deg or so) EDS quantitative analysis shows a decrease
of O2%. How this is possible? I expected to see O2% goes up
with the tilt as electron will see more of the surface than
bulk.

Could someone explain why?

Thanks

Fran.


---------------------------------------------------------
Get Your *Web-Based* Free Email at http://www.hotmail.com
---------------------------------------------------------




From: richard.lander-at-stonebow.otago.ac.nz (Richard Lander)
Date: Tue, 4 Mar 1997 12:52:34 +1200
Subject: Thanks Re: Disposable gloves and Multi user Security.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear all,

Many thanks for the many recent replies to my earlier queries on Disposable
gloves and multi user security. We now have a few more ideas on where to
go regarding security in our Unit.
The gloves issue is very complex too, mainly due the wide variety of
chemicals and conditions they are used for in the lab!

Thanks again,

Rich.

-----------------------------------------------------------------------
Richard Lander, NZCS
South Campus Electron Microscope Unit
c/- Pathology Department
Otago Medical School
P.O. Box 913
Dunedin
New Zealand.
Tel. National 03 479 7301 Fax. National 03 479 7254

"Southernmost EM Unit in the World!"
------------------------------------------------------------------------






From: Gang REN :      ren-at-image.blem.ac.cn
Date: Tue, 4 Mar 1997 10:29:00 -0800
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Unsubscribe me please

--
+--------------------------------------------------------------------+
| Gang REN, Ph.D Candidate Email: ren-at-image.blem.ac.cn |
| Beijing Lab. of Electron Microscopy ren-at-indy1.MSE.CWRU.Edu |
| Center for Condensed Matter Physics Tel: (+86-10) 62568304 |
| Chinese Academy of Sciences Fax: (+86-010) 62561422 |
| P.O. Box 2724, Beijing 100080 URL:http://image.blem.ac.cn/ |
| P.R. China ren_home/ren.html |
+--------------------------------------------------------------------+






From: John Best :      jbest-at-vicon.net
Date: Tue, 04 Mar 1997 08:32:10 -0800
Subject: Re: EDS related question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

fransisco black wrote:
} Dear Microscopists:
} I would like to get an explanation for the following observation
} of quantitative analysis by EDS.
}
} I am looking at a Si wafer with 100nm of SiO2 layer on top.
} from a SEM. Electron beam (~8 kev) hit the surface along the
} surface normal (0 deg tilt). As I increase the sample tilt
} (upto 45 deg or so) EDS quantitative analysis shows a decrease
} of O2%. How this is possible? I expected to see O2% goes up
} with the tilt as electron will see more of the surface than
} bulk.
} Could someone explain why?
} Thanks
} Fran.

Fran,
My initial guess would be the change in the sample-detector
geometry. Although one would expect the O2 counts to go up, the detector
may no longer be in as good a position (relative to the sample) to pick
them up. You might simply adjust the insertion of the EDS detector to
compensate.

One could use an interpolative process: adjust the tilt a bit, then
adjust the detector insertion to maximize counts. Then adjust the tilt a
little more, and readjust the detector insertion.

You might get to an optimum geometry faster if you do a rough calculation
of the angle of incidence, reflection and best detector insertion to
intersect.

Regards,
John Best -- ELMDAS Co.
http://www.vicon.net/~jbest




From: William Tivol :      tivol-at-wadsworth.org
Date: Tue, 4 Mar 1997 09:05:35 -0500 (EST)
Subject: Re: EDS related question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} I am looking at a Si wafer with 100nm of SiO2 layer on top.
} from a SEM. Electron beam (~8 kev) hit the surface along the
} surface normal (0 deg tilt). As I increase the sample tilt
} (upto 45 deg or so) EDS quantitative analysis shows a decrease
} of O2%. How this is possible? I expected to see O2% goes up
} with the tilt as electron will see more of the surface than
} bulk.
}
Dear Fran,
What is the geometry (take-off angle, direction of tilt, etc.)?
It may be that some O x-rays are taking a longer path and are absorbed,
or the Si x-rays can take a shorter path and be absorbed less. Have
you used David Joy's Monte Carlo program to calculate the volume in
which the x-rays are generated?
Yours,
Bill Tivol





From: Woody.N.White%650 :      Woody.N.White-at-mcdermott.com
Date: 3/3/97 6:20 PM
Subject: EDS related question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


All else being the same, I (as it sounds you would) would expect an increase
in
the relative intensity of the oxygen (raw counts). Do you see this on the
graphical result? If so, it is most likely that the ZAF/PhiRhoZ/etc. quant.

algorithm is the source of the problem. Woody
______________________________ Reply Separator
_________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Dear Microscopists:

I would like to get an explanation for the following observation
of quantitative analysis by EDS.

I am looking at a Si wafer with 100nm of SiO2 layer on top.
from a SEM. Electron beam (~8 kev) hit the surface along the
surface normal (0 deg tilt). As I increase the sample tilt
(upto 45 deg or so) EDS quantitative analysis shows a decrease
of O2%. How this is possible? I expected to see O2% goes up
with the tilt as electron will see more of the surface than
bulk.

Could someone explain why?

Thanks

Fran.


---------------------------------------------------------
Get Your *Web-Based* Free Email at http://www.hotmail.com
---------------------------------------------------------




From: wise-at-vaxa.cis.uwosh.edu
Date: Tue, 04 Mar 1997 09:13:27 +0000
Subject: Supplies and teaching

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To all,

I teach electron microscopy (both TEM and SEM) as a one semester
course. The first couple of times I had 3 to 8 students but last fall I
had 14. Students used photographic film and paper out of a common supply.
It seemed that every time I turned around, we were out of paper and film or
some unidentified person had exposed a portion of our common supply to
light. In order to better monitor usage and I am considering giving each
student his/her own allotment of paper and film at the beginning of the
semester. I require them to turn in 10 TEMs and 10 SEMs at the end of the
semester.
What do those of you who teach EM (and old fashioned darkroom
techniques) require for your students to turn in and how do you budget
and/or distribute supplies?

Bob


Robert R. Wise
Plant Physiologist and Director, UWO Electron Microscope Facility
Department of Biology
University of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404 tel
(414) 424-1101 fax
wise-at-uwosh.edu






From: gfritz-at-noran.com (Greg Fritz)
Date: Tue, 4 Mar 1997 09:49:58 -0600
Subject: Re: EDS related question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Dear Microscopists:
}
} I would like to get an explanation for the following observation
} of quantitative analysis by EDS.
}
} I am looking at a Si wafer with 100nm of SiO2 layer on top.
} from a SEM. Electron beam (~8 kev) hit the surface along the
} surface normal (0 deg tilt). As I increase the sample tilt
} (upto 45 deg or so) EDS quantitative analysis shows a decrease
} of O2%. How this is possible? I expected to see O2% goes up
} with the tilt as electron will see more of the surface than
} bulk.
}
} Could someone explain why?
}
} Thanks
}
} Fran.
}

The problem observed here is most likely related to simple rules of
geometry. If the EDS detector is not perpendicular to the tilt axis you
will find that the path through the sample that the xrays must use to get to
the EDS detector becomes longer as the tilt axis increases. (Lower observed
take off angle). This results in more absorption. The lower energy peaks
will be abosrbed at a greater rate than the higher energy lines. If you are
not taking the azimuth angle into your take off angle correction you should.

If you want to limit your xray excitation volume to give you a better idea
of the film try also using a lower accelerating voltage. Normally an
excitation energy of 2.5 times your xray energy should work.

Greg Fritz
NORAN Instruments





From: I.Montgomery-at-biomed.gla.ac.uk (Ian Montgomery)
Date: Tue, 4 Mar 1997 17:07:30 +0000
Subject: Brendel tissue slicer.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Has anyone used the Brendel tissue slicer and what do you think of it.
Ian.






From: Steven D. Majewski :      sdm7g-at-Virginia.EDU
Date: Tue, 4 Mar 1997 17:35:54 -0500 (EST)
Subject: Re: MSA/MAS spectrum file format

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


BTW: I have a DTSA binary -} EMMFF ascii file converter written in
Python if anyone needs one. It actually writes a somewhat loose
version -- the spec was ambiguous in several places and I loosened
some restrictions intentionally for my own purposes. It's written
in Python, and it's easily modifiable to your own customizations.
( Specifically, I use longer user defined keyword fields to save
the original DTSA header information using the names in the DTSA
file interface definition. The DTSA EMMFF plugin seems to read
these files back in properly -- the standard tags that is. It
ignores the original saved DTSA information. )

I wrote this code before there was a plug-in for DTSA, however I
still require it to do batch conversions of folders full of files.

The current version is specifically for the Mac. The initial version
was originally run on unix, and most of it should be portable. The
main mac specific feature is that it supports drop launching of a
whole folder of DTSA files. When I have the final, portable version
( hopefully with a stand-alone Mac version that doesn't require
Python installed. ) it will be posted on the Web. If anyone needs
it right away, I can email you a copy of the sources.


---| Steven D. Majewski (804-982-0831) {sdm7g-at-Virginia.EDU} |---
---| Department of Molecular Physiology and Biological Physics |---
---| University of Virginia Health Sciences Center |---
---| P.O. Box 10011 Charlottesville, VA 22906-0011 |---
By doing just a little every day, you can gradually
let the task completely overwhelm you.





From: Leah Dobbs :      ldobbs-at-itis.com
Date: Tue, 4 Mar 1997 19:16:38 -0000
Subject: Re: Supplies and teaching

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am a student of EM at Madison Area Tech. We had to buy our own paper. I
am now a lot better at making prints !

Leah



----------
} From: wise-at-vaxa.cis.uwosh.edu
} To: Microscopy-at-Sparc5.Microscopy.Com
} Subject: Supplies and teaching
} Date: Tuesday, March 04, 1997 9:13 AM
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} To all,
}
} I teach electron microscopy (both TEM and SEM) as a one semester
} course. The first couple of times I had 3 to 8 students but last fall I
} had 14. Students used photographic film and paper out of a common
supply.
} It seemed that every time I turned around, we were out of paper and film
or
} some unidentified person had exposed a portion of our common supply to
} light. In order to better monitor usage and I am considering giving each
} student his/her own allotment of paper and film at the beginning of the
} semester. I require them to turn in 10 TEMs and 10 SEMs at the end of
the
} semester.
} What do those of you who teach EM (and old fashioned darkroom
} techniques) require for your students to turn in and how do you budget
} and/or distribute supplies?
}
} Bob
}
}
} Robert R. Wise
} Plant Physiologist and Director, UWO Electron Microscope Facility
} Department of Biology
} University of Wisconsin Oshkosh
} Oshkosh, WI 54901
} (414) 424-3404 tel
} (414) 424-1101 fax
} wise-at-uwosh.edu
}
}




From: richard.lander-at-stonebow.otago.ac.nz (Richard Lander)
Date: Wed, 5 Mar 1997 15:57:49 +1200
Subject: Fluoronanogold for EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message on behalf of Allan Mitchell;

Greetings,
I have a user of our EM Unit who wishes to use Fluoronanogold (Nanoprobes)
with Silver enhancement for pre-embedding immuno EM. We will be using
frozen 8 micron sections which will then be PLP fixed (parform + Lysine +
periodate) prior to labelling.
After labelling, we want to post-fix them in OsO4 (as membranes are
important), dehydrate and embed in normal epoxy resin. I have heard that
the Flouronanogold is not stable at 60oC, the normal curing temperature of
the resin.

Has anybody had experience with pre-embedding with Fluoronanogold?
Will Fluoronanogold survive the epoxy resin curing process at 60oC?
If it wont survive, which resin and curing procedure would be the choice to
use in this case? (Quetol 651, Lowicryl?)

Many thanks,

Allan Mitchell.






From: le_thiec-at-nancy.inra.fr (Didier Le Thiec)
Date: Wed, 5 Mar 1997 10:30:25 -0500
Subject: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear all,

I am interested to replace our SEM by ESEM or low vacuum SEM, but I
would like to have more information about EDX microanalysis with these SEM
(what are the quantitative and qualitative differences when I change the
pressure? what's happen with cryo system .....?).
Thanks in advance for your help (comments, references ...)

Best regards to all
Didier

--------------------------------------------
Dr. Didier Le Thiec
I.N.R.A. Centre de Recherches Forestieres
Unite d'Ecophysiologie Forestiere
Laboratoire de Pollution Atmospherique
54280 Champenoux - France

Tel : 33 (0)3 83 39 40 98
Fax : 33 (0)3 83 39 40 69
E-mail : le_thiec-at-nancy.inra.fr
-------------------------------------------






From: MEV H VAN DER MERWE :      HVDM-at-op1.up.ac.za
Date: Wed, 5 Mar 1997 14:43:45
Subject: Immuno-Bed TM kit

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Can anyone please tell me all the uses of the Immuno-Bed Kit?

Thanks

Hildagonda van der Merwe


University of Pretoria
Faculty of Veterinary Science
Dept. of Pathology
Onderstepoort
0110

e-mail: HvdM-at-op1.up.ac.za




From: John Best :      jbest-at-vicon.net
Date: Wed, 05 Mar 1997 09:03:58 -0800
Subject: Science Teacher Needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello All............

First, my apologies to the Microscopy listserver subscribers if this posting isn't appropriate. We need a
first rate science teacher with experience, and my colleagues here at the microscopy listserver sprung to mind
immediately.

Secondly, If your not interested in accepting a position as a high school science teacher, and know of a
listserver that may produce some interested candidates, please forward this request. Thank You.

And finally: The Juniata Valley School District is seeking candidates for the position of Science Teacher
beginning in the fall of 1997. Responsibilities are: fundamentals of science and physical sciences for
students in grades 7-12.

We're looking for candidates who bring a number of years of experience from industry or a university research
environment to the classroom. Pennsylvania teaching certification is required, or must be readily obtainable.
The ability to motivate and educate "typical" high school students is a must.

With a BS the starting salarary is $28,041, an MS starts at $28,712. Starting salaries are dependant on
teaching experience within the Pennsylvania public school system.

The Juniata Valley School District has approximately 500 students in grades 7-12 and is situated in a beautiful
rural area approximately 25 miles southwest of the main campus of The Pennsylvania State University. JVSD is
an equal opportunity employer.

Please forward your resume to our superintendant, Dr. David Leckvarcik at The Juniata Valley School District,
P.O. Box 318, Alexandria, PA 16611. If you would like me to follow up on your resume, please forward it to me
at: jbest-at-vicon.net.

Thank You.
John Best - JVSD School Director.




From: David Knecht :      knecht-at-uconnvm.uconn.edu
Date: Wed, 5 Mar 1997 08:22:34 -0600
Subject: Temperature control for stage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am wondering what solutions people have found to the stage
temperature/gas/humidity control problem for tissue culture microscopy of
live cells. I have tried an open chamber approach (20/20 Technologies) and
evaporation onto the condenser or lid was a major problem as was even
heating. Are there any commercial stage temperature controllers that rely
on hooding the microscope or stage that work well? Has anyone constructed
something simple that can be reproduced easily? What are the down sides of
this approach? Has anyone used the new Bioptechs system and does it
actually do the job and solve any or all of these problems? Thanks- Dave

Dr. David Knecht
Department of Molecular and Cell Biology
University of Connecticut
U-125
Storrs, CT 06269
Knecht-at-uconnvm.uconn.edu






From: ALLEM, Rafik :      allem-at-paprican.ca
Date: Wed, 5 Mar 1997 09:42:17 EST5EDT
Subject: SEM_stereo_microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for a PC based software (preferably on Windows) that can
perform profilometry type measurements from an SEM pair of images
(grabbed into 'TIF' files).
Measurements should include:
- Line profiles
- Height measurements
- Contour maps.

Any help is appreciated.

Regards,

Dr. Rafik Allem
Pulp and Paper Research Institute of Canada.
570, St. John's Boulevard, Pointe-Claire,
Qc, H9R 3J9, Canada.
tel: (514)630-4101 ext. 2661, fax:(514)630-4134
e-mail: allem-at-paprican.ca






From: Norma Procyshyn :      NORMA-at-MINMET.Lan.McGill.CA
Date: Wed, 5 Mar 1997 09:49:14 EST5EDT
Subject: Re: McGill seminar

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Department of Mining and Metallurgical Engineering
Professional Development Seminars

"MATERIALS CHARACTERIZATION FOR THE METAL, MINERAL AND
MICROELECTRONICS INDUSTRY"
May 12-15, 1997

Seminar Leader: Neil Rowlands

This course is intended to benefit users who are concerned with the
analysis of fine structures, fine particles and may also have the
need to image surfaces down to the atomic level. These techniques
will be of special interest to those involved in the
microelectronics, polymer science, and ceramics industries, as well
as those involved in metallurgical analysis.

The course will be of a practical nature and will be of interest to
operators and managers of analytical laboratories R&D engineers and
applied scientists. Examples of the applications of each technique
will be presented and on the last day of the course there will be an
optional visit to the Materials Technology Laboratory, CANMET in
Ottawa to see many of these techniques in operation. Additional
facilities (e.g. TEM, AFM) are available on the McGill campus.

For more information please contact

Norma Procyshyn
Dept. Mining & Metallurgical Eng.
McGill University
2020 University Street, Box 102
Montreal, Quebec H3A 2A5
Telephone 514-398-4383
Fax 514-398-6044
e-mail norma-at-minmet.lan.mcgill.ca







From: nano-at-ns1.lihti.org (Nanoprobes, Inc.)
Date: Wed, 5 Mar 1997 15:17:30 -0500
Subject: Re: Fluoronanogold for EM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Allan:

I already replied privately before I got down our in-box to the MSA
posting. For anyone interested in this, here are our ruminations...

Generally, we recommend low-temperature polymerization resins such as
Lowicryl, LR White or any similar medium for embedding Nanogold=81 or
=46luoroNanogold-labeled specimens. This is because some early experiments
with Nanogold=81-labeled specimens suggested that the higher-temperature
embedding (60=B0C) might cause the gold particles to be displaced from their
binding sites. However, subsequent experiments with Nanogold=81 in solution
showed that it could be heated even up to 100=B0C for 1 h with minimal
decrease in optical density; generally, avoiding low pH values (below 7) or
high ionic strengths (0.2 M NaCl or higher) helps ensure its stability.
Because it is smaller than most colloidal gold and has no tendency to stick
electrostatically to proteins or cell components, it may be more free to
move: fixing with glutaraldehyde helps counteract this, and with Nanogold=81
we also recommend silver enhancement before embedding if it is practical.

There is a section on this in Hainfeld and Furuya's chapter on the silver
enhancement of Nanogold=81 and undecagold in M. A. Hayat's recent book,
"Immunogold-Silver Staining: Principles, Methods and Applications" (CRC
Press, Boca Raton, 1995; pp. 71-96. Check pages 92-92 for the effects of
heating Nanogold=81. From this section, heating at 60=B0C for 250 minutes
resulted in a reduction of the optical density to 80% of its initial value
- suggesting that Nanogold=81 can survive most 60=B0C embedding procedures.

The gold particle in FluoroNanogold is Nanogold, and exactly the same
applies to this probe. However, silver enhancement quickly removes the
fluorescence, so only do the pre-embedding silver enhancement if you have
completed the fluorescence microscopy.

We keep a list of answers and suggestions to frequently-asked questions in
the Technical Help section of our web site:

http://www.nanoprobes.com/Tech.html

All the suggestions made there about Nanogold=81 use and stability also appl=
y
to Fluoronanogold, since these probes use the same gold particle.

Hope this is helpful,

Rick Powell



******************************************************************
* NANOPROBES, Incorporated | Tel: (516) 444-8815 *
* 25 East Loop Road, Suite 124 | Fax: (516) 444-8816 *
* Stony Brook, NY 11790-3350, USA | nano-at-mail.lihti.org *
* *
* NOW EASY TO FIND ON THE WEB: http://www.nanoprobes.com *
******************************************************************






From: Goodhouse, Joseph :      jgoodhouse-at-molecular.princeton.edu
Date: Wed, 05 Mar 97 10:09:00 PST
Subject: Re- Fluoro-Nanogold

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In response to Allan Mitchell's inquiry.
If you are speaking of whether the Fluorescent signal of
Fluoro-Nanogoold survives the answer is no, but not because of the 60 degree
heat. The fluorescence is quenched or covered up by the silver enhancement
process. If you want to see the label prior to silver enhancement that is
no problem. I often do this prior to running the silver enhancement to see
that the reagent has gotten in to the sample and gives the expected signal.
Your other choice of resins are lowicryl which you can polymerize at -35
degrees under UV. The FLUOR tag survives this just fine. In regard to
osmication, this has to be done with 0.1% osmium at 4 degrees or on ice for
about 30 - 45 minutes. Standard osmication procedures will reduce the
silver shell that forms around the nanogold particle. This appears as a much
less electron dense cloud or ghost. I have used these reagents with spurr,
epox 812, lowicryl and LR White without any problem.

Joe Goodhouse
Confocal / E.M. Lab
Molecular Biology
Princeton University




From: Michael D Standing :      MDStandi-at-bioag.byu.edu
Date: Wed, 05 Mar 1997 09:00:55 -0600 (MDT)
Subject: Re: Supplies and teaching

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} From: wise-at-vaxa.cis.uwosh.edu
} Date: Tue, 04 Mar 1997 09:13:27 +0000
} Subject: Supplies and teaching
} To: Microscopy-at-Sparc5.Microscopy.Com

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} To all,
}
} I teach electron microscopy (both TEM and SEM) as a one semester
} course. The first couple of times I had 3 to 8 students but last fall I
} had 14. Students used photographic film and paper out of a common supply.
} It seemed that every time I turned around, we were out of paper and film or
} some unidentified person had exposed a portion of our common supply to
} light. In order to better monitor usage and I am considering giving each
} student his/her own allotment of paper and film at the beginning of the
} semester. I require them to turn in 10 TEMs and 10 SEMs at the end of the
} semester.
} What do those of you who teach EM (and old fashioned darkroom
} techniques) require for your students to turn in and how do you budget
} and/or distribute supplies?
}
} Bob
}
}
} Robert R. Wise
} Plant Physiologist and Director, UWO Electron Microscope Facility
} Department of Biology
} University of Wisconsin Oshkosh
} Oshkosh, WI 54901
} (414) 424-3404 tel
} (414) 424-1101 fax
} wise-at-uwosh.edu
}
}
}
Bob,

We here at BYU teach a course in SEM and in TEM. We require our
students to submit two types of assignments to complete the course.
First is a portfolio which may or may not use traditional darkroom
techniques (digitized, laser printed images are an option from our
SEMs). We also have the students turn in a poster representing a
project they have worked on all semester. The poster must have at
least 5 high quality photographs which the student has produced, as well as any text and graphics the
student wants to display. The student purchases all of the
photographic materials they use. We do keep a stock supply in
the lab for the students to purchase from.
Students are responsible for all of their own photographic supplies
so if thier paper becomes light struck, that is their problem. This
has worked quite successfully for the past several years.

I hope this helps

Michael D. Standing
e-mail: MDStandi-at-bioag.byu.edu
Phone: (801)378-4011




From: Tom Phillips :      tphillips-at-biosci.mbp.missouri.edu
Date: Wed, 5 Mar 1997 10:11:14 -0500
Subject: Ilford photo processor problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have noted several users of this listserver refer to the Ilford RC2150
print processor. We have used one for several years without trouble but
Ilford has apparently switched to a new "environment friendly" cleaning
solution that is unable to remove the deposits from the rollers. All of
our prints have terrible streaks of silver residue. Has anybody else had
this problem - more importantly, has any been else solved it? TIA


Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility
3 Tucker Hall
University of Missouri
Columbia, MO 65211
(573)-882-4712 (voice)
(573)-882-0123 (fax)






From: EmLab :      EMLAB-at-OPUS.MCO.EDU
Date: Wed, 05 Mar 1997 11:50:16 -0500 (EST)
Subject: Re: Supplies and teaching

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Bob,

Way back when, the TEM class I took, it was required for me to
purschase my own film, paper, tweezers, embedding mold and a text book. All
this was setup by the proffessor and sold thru the bookstore on campus.
Yes it was/is expensive but it taught me to be conservative with these
items. Most of the prints I made for the class were 4x5. The class was ten
weeks long and I did not run out of supplies.

Hopes this helps,

Ed Calomeni
Dept. Pathology
Medical College of Ohio
Toledo, Ohio
emlab-at-opus.mco.edu




From: Regina Messer :      messer52-at-eng.uab.edu
Date: Wed, 5 Mar 1997 11:38:10 -0600
Subject: LM-Stains for metals in fibroblast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am trying to use stains discribed in
Histochemistry: Theoretical and Applied Volume 2 by AGE Pearse
to locate metallic ions with in tissue. Any suggestions for other
methods will be appreciated.
The following are the stains, the metal I'm trying to detect, and the
problem I am having. I trying to stain cells not sections.

Napthochrome Green B Beryllium (1)unable to find acridine red
Aldridge says is was discontinued 12 years ago

Rubeanic Acid Nickel (1) the stain is very light and is not
staining cells

Chrome Azurol S Chromium (1)library is unable to get referenced
paper and therefore I am unable
to find protocol

If anyone has any information that might help please responed.


Regina Messer
Graduate Student
Biomedical Engineering
UAB




From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Wed, 05 Mar 1997 18:05:04 -0800
Subject: Re: stains for metals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

For staining metals look in "Histopathologic Technique and
Practical Histochemistry" by Ralph Lillie and Harold Fullmer; McGraw
Hill, 1976. Out of print for years but probable available in your library
or via interlibrary loan.
Also "Selected Histopathological Methods" or something like that
by Thompson is a massive tome with just about everything in it.
Call or e-mail if necessary.

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (908)-235-4583; fax -4029 e-mail: mcauliff-at-.umdnj.edu
***************************************************************




From: MVARGAS-at-dentistry-po.dentistry.uiowa.edu
Date: Tue, 4 Mar 97 23:10 CST
Subject: Collagen stain

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Does anybody know a good stain or protocol to stain type I collagen
for High resolution SEM

marcos vargas




From: Warren Straszheim :      wes-at-ameslab.gov
Date: Wed, 05 Mar 1997 17:42:13 -0500 (CDT)
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

We have observed that spatial x-ray resolution suffers, even at 40 Pa with a
25 mm working distance. We can get decent x-ray maps, but quantitative
analyses can easily pick up tenths of a percent of an element from a nearby
phase. We have yet to quantify the effects for ourselves. I recall there
were a number of papers at the Minneapolis meeting on this effect. One has
to be careful.

At 10:30 AM 3/5/97 -0500, you wrote:

} Dear all,
}
} I am interested to replace our SEM by ESEM or low vacuum SEM, but I
} would like to have more information about EDX microanalysis with these SEM
} (what are the quantitative and qualitative differences when I change the
} pressure? what's happen with cryo system .....?).
} Thanks in advance for your help (comments, references ...)
}
} Best regards to all
} Didier
}
} --------------------------------------------
} Dr. Didier Le Thiec
----------------------------------------------------
Warren E. Straszheim
270 Metals Development, Ames Lab/ISU, Ames IA, 50011
Phone: 515-294-8187 FAX: 515-294-3091

E-Mail: wes-at-ameslab.gov (or: wesaia-at-iastate.edu)
http://www.public.iastate.edu/~iprt_info/cfce/ (re: coal)
http://www.public.iastate.edu/~wesaia/marl/ (re: SEM)

coal characterization and processing
electron microscopy, x-ray analysis, image analysis
computer applications





From: Peter Jordan :      emsi-at-pe.net
Date: Wed, 05 Mar 1997 19:14:33 -0800
Subject: Looking for used SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello:
I am looking for a used SEM with low Voltage capability, about 1 KV. Let
me know if you know of anybody who would like to sell. Please E-mail or
call me at 909 694-1839.
Thanx, Peter Jordan




From: Tony Bruton :      bruton-at-EMU.UNP.AC.ZA
Date: Thu, 06 Mar 1997 08:31:54 +0200
Subject: Re: SEM - EDX -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Warren

Can you tell us about the instruments that you are using ?

Tony Bruton
Centre for EM
University of Natal
Pietermaritzburg
South Africa





From: SONEJA A K :      soneja-at-giasbma.vsnl.net.in
Date: Thu, 6 Mar 1997 13:27:59 +0530 (IST)
Subject: Re: SEM - EDX -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello guys.

I would be highlyobliged if someone could guide me to the sources for the
following image analysis solutions.

1} Visual inpection systems for textile fibres/i textile industrry.
2} Image analysis software for textile fibres,sperm cell analysis.

Names of organisations dealing in these areas with their email/home
page/fax numbers solicited.

Thanks.

Best regards,
Anish

*************************************************************************
For further details please contact:
Soneja A.K.
Director
METZER BIOMEDICAL & ELECTRONICS PVT.LTD.
327 Wadala Udyog Bhavan,Wadala,MUMBAI(BOMBAY )400 031.INDIA
Tel:91 22 4145057/4165650
Fax 91 22 4168757

Email:soneja-at-giasbma.vsnl.net.in
*************************************************************************





From: Joergen Bilde-Soerensen 5802 :      j.bilde-at-Risoe.DK
Date: Thu, 06 Mar 1997 09:35:19 +0100
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Didier Le Thiec wrote:
} I am interested to replace our SEM by ESEM or low vacuum SEM, but I
} would like to have more information about EDX microanalysis with
} these SEM (what are the quantitative and qualitative differences when
} I change the pressure? what's happen with cryo system.....?)

Dear Didier,

The presence of gas in the specimen chamber influences the spatial
resolution of EDX microanalysis in ESEM and LVSEM. The primary
electrons will scatter on the gas and give rise to a skirt of
scattered electrons that will excite X-rays at places pretty far away
from the point you want to analyze. The scattered intensity is
approximately given by Is/Io = 1 - exp(-psL/kT) where p is the
pressure, s the total scattering cross section for electron scattering
on the gas used, L the distance between the last pressure limiting
aperture and the sample, k the Boltzmann constant and T the absolute
temperature. Examples of skirt shapes are given in:

D. A. Moncrieff et al., J. Phys. D: Appl. Phys. vol. 12 (1979)
481-88.
D. C. Joy, Microscopy and Microanalysis ' 96, Proc. Annual
Meeting MSA, Minneapolis, Minnesota, 11 - 15 August 1996

It is to a large degree possible to correct for the beam skirt
effects:

1) You can extrapolate from spectral measurements made at several
different pressures to the result that would have been found without
scattering provided that the measurements are made in the single
scattering regime (e.g. pL { approx. 1.6 Pa.m for measurements in
water vapour).
2) If there is plural scattering, you can take two
spectra, one with a fine needle (of the kind used for field ion
microscopy or scanning tunneling microscopy) inserted over the point
of interest, and the other with the needle slightly retracted.
Subtraction of the first from the second spectrum will approximately
give the spectrum from the point of interest.

Neither method will give you as exact an analysis as you will get
under high vacuum, but you can get rid of most of the skirt effects.
The pressure variation method in particular yields pretty good results
if carefully performed. The methods are described in:

J. B. Bilde-Soerensen and C. C. Appel, Proc. 48th Annual Meeting of the
Scandinavian Society for Electron Microscopy, Aarhus, 2 - 5 June 1996,
pp. 4 - 5.
J. B. Bilde-Soerensen and C. C. Appel, Proc. 11th European
Congress on Microscopy EUREM' 96, Dublin, 26-30 August 1996. Session
T6.

ESEMs and LVSEMs can also be operated as high vacuum SEMs - so you
still have the possibility of going to high vacuum for microanalysis.

Best wishes,
Jorgen.
-at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at--at-
J. B. Bilde-Soerensen
Materials Research Department
Risoe National Laboratory
DK-4000 Roskilde
Denmark

e-mail: j.bilde-at-risoe.dk
phone: +45 46 77 58 02 (direct)
phone: +45 46 77 46 77 (switchboard)
fax: +45 46 35 11 73
website: http://www.risoe.dk




From: chris gilpin :      CGILPIN-at-fs1.sem.man.ac.uk
Date: Thu, 6 Mar 1997 12:39:48 BST
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

We have published a couple of papers on ESEM EDX. Might be worth a
look. :)
I would like to add that if you wish to look at hydrated samples then
the chamber pressures required are such that scattering of the
primary beam results in a probe at the sample of at least 1mm
diameter!!
Also correction methods based on taking spectra at different
pressures are fraught with difficulty if you need to avoid drying out
your sample.
Also be aware that with low Z detectors a contribution from the
chamber gas will be encountered.

Chris



} I am interested to replace our SEM by ESEM or low vacuum SEM, but I
} would like to have more information about EDX microanalysis with these SEM
} (what are the quantitative and qualitative differences when I change the
} pressure? what's happen with cryo system .....?).
} Thanks in advance for your help (comments, references ...)
}
} Chris Gilpin
Biological Sciences Electron Microscope Unit
G452 Stopford Building
Oxford Road
Manchester
M13 9PT
phone +44 161 275 5170
fax +44 161 275 5171




From: ebs-at-ebsciences.com
Date: Thu, 06 Mar 1997 08:23:45 EST
Subject: Immuno-Bed kit

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear fellow microscopists,

At 02:43 PM 3/5/97, Hildagonda van der Merwe wrote:
} Can anyone please tell me all the uses of the Immuno-Bed Kit?

Immunobed is an embedding medium for immunohistochemistry which allows the
rapid penetration of large immunoglobulins through the section for
demonstration of antigenic sites.

For more detailed information, please request our A2050 data sheet.

Best regards,
Steven E. Slap, Vice-President
********************************
Energy Beam Sciences, Inc.
Adding Brilliance To Your Vision
ebs-at-ebsciences.com
http://www.ebsciences.com/
********************************





From: Andrey L. Chuvilin :      dusha-at-catalysis.nsk.su
Date: Thu, 6 Mar 1997 19:37:50 +0700 (GMT)
Subject: Re: DIGITAL IMAGING

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I wish to thank those people (Scott Whittaker, Gary Chinga, Christian Mellen,
Melvyn Dickson, John Mardinly, Hans Tietz) who replied me, through there were
a few. (Sprinkling water seems to be much more interesting to talk
about.:)
Short summary is:
The main software for image analysis is NIH (FTP from
zippy.nimh.nih.gov), through analySIS also exists and Photoshop,
Paintshop, Coreldraw and Wincatpro can be used to manipulate images.

CD-Rs mainly used to store, transport and distribute image files.

1200dpi LP (mainly Lexmark) used for draft and dye-sub printers for
quality prints. So one will need both...

Useful references obtained:
http://www.biotech.ufl.edu/~emcl/ The home of " Tips & Tricks "
site supported by Scott Whittaker - many discussions from
this Listserver and Confocal Listserver archived there.
Extremely useful!

http://www.soft-imaging-web.de home of Soft-Imaging Software
information about analiSIS software package, etc.

Stewart, M.G. and Davies, H.A. (1995) Digital image processing (DIP) in the
TEM: is it viable in biological morphometry?. Eur Microsc Anal (1995) 35,
pp 21-23.

Special opinion: "Andrej! Do not change!..."

Full text of replies can be obtained on request.
A. Chuvilin






From: Jitu Shah :      JSS-at-siva.bris.ac.uk (by way of Nestor J. Zaluzec)
Date: Thu, 6 Mar 1997 07:51:48 -0500
Subject: SEM-EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Dr. Didier,

I saw your concern about the EDX at higher pressures. As is suggested by
others
there are inaccuracies in the quantitative estimation. This is because the
incident beam coming through the gaseous region spreads before it hits the
specimen surface. The electrons outside the original gaussian profile of the
beam also create x-rays. You are probably aware of David Joy's work which was
presented at MSA meeting in the US. We are enormously interested in this
problem
and one of my students is doing Monte carlo calculations on this very problem.
You also inquired about cryoSEM. Normally, you should not have any problem
about
the beam scattering as the equilibrium vapour pressure is below the chamber
pressure. So you can assume that all of your x-rays are coming from the
original
focussed beam spot. This of course assumes that you are not using a very
intense
beam which evaporates water or other constituents in your specimen and
upsetting
the inherent composition of the specimen. Incidentally,at Bristol we convert
conventional SEMs in to High pressure SEMs exceeding the performance of the
commercially avilable instruments. If you need any help on this please do not
hesitate to get in touch with me

Yours sincerely

Jitu Shah

Dr.Jitu Shah
H.H. Wills Physics Laboratory,
University of Bristol,
Royal Fort, Tyndall Avenue,
Bristol BS8 1TL. UK
email: jss-at-siva.bristol.ac.uk
Tel: 44 117 9288719
Fax: 44 117 9255624






From: MARK DARUS (216) 266-2895 GENERAL ELECTRIC CO. :      darus-at-cle.dnet.ge.com
Date: Thu, 6 Mar 97 08:34:34 EST
Subject: Al peak

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

When I perform an EDX analysis on uncoated fused silica, 10 20 or 30 KeV,
and the sample begins to charge, often I get a small aluminum peak, and it
doesn't go away as I move around the sample, it comes back. The amount of
aluminum in the quartz is typically less than 20 ppm, so I don't think that
the Al peak is coming from my sample.
Could it be that the Al peak is coming from my sample stage or walls of
the sample chamber, due to any effect caused by the charge that has developed
on the uncoated quartz?
Also, for EDX work, what are the negative effects that sample charging
has on an analysis? My work is done on an Amray 1600 with a Noran detector,
one that has two windows, Be and the ultra thin window.

Thanks, Mark Darus
Darus-at-cle.dnet.ge.com




From: Zhiyu Wang :      wangz-at-pulsar.cs.wku.edu
Date: Thu, 6 Mar 1997 09:08:24 -0600 (CST)
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear Bilder:

Recently I got a problem in LVSEM-EDS system. The SEM is JSM-5400LV with
Kevex EDS. We use a thin B-window with expect to determine elements down
to Z=5. The SEM does not have pre-vacuum chamber which means whenever we
change sample, the chamber and whole column are exposured to atmosphere.
The window has been used for 10 months with good performance. Therefore,
recently once we change sample as usual, the window got brocken. We have
no idea what is the reason.

My question is: How thick window should I use to prevent the window from
brocken, and what is the common technical hints to prevent the window from
brocken? Does the frequently presure change on window surface (when
change sample) affect the life time of window?


Thank you very much


Zhiyu Wang
Electron Microscope Facility
Western Kentucky University
Bowling Green KY 42103
USA
wangz-at-pulsar.cs.wku.edu





From: tong-at-cebaf.gov (Tong Wang)
Date: Thu, 06 Mar 1997 12:08:39 -0500
Subject: Carbon analysis quantification using EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am trying to do trace Carbon analysis in Nb material. The Be window of my
EDX can be opened in order to detect C, N, O, but the quantification seems
impossible. Anybody has similar experience?

Any response is appreciated!

Best regards.

Tong






From: cheng-at-ems.psu.edu (Shang-Cong Cheng)
Date: Thu, 6 Mar 1997 13:12:31 -0500
Subject: Carbon analysis quantification using EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

subscribe

Shangcong Cheng
196 MRI Building, PSU Research Park, University Park, PA 16802
Tel: (814) 863-3767 Fax: 863-0637 Email: cheng-at-ems.psu.edu






From: Giles John E Jr :      giles_john_e_jr-at-space.honeywell.com
Date: 6 Mar 1997 13:24:28 -0500
Subject: SEM Courses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hello Everyone,

I have been tasked with bringing another technician into our SEM lab and would
like to get info on the courses which are available for a person who is
familiar with light microscopy but knows almost nothing about a SEM.

The desired course would be 5 days or less and be applicable to the level of
knowledge described above. We deal with strictly materials analysis, so
biological based courses are not applicable. The desired outcome would be
someone who has a rudimentary knowledge of SEM theory and capable of taking
basic SEM pictures and acquiring EDS spectra on non-challenging samples. We
will be doing some training of the new tech, but due to heavy workloads, we
can't devote the time to the training that we would normally do.

Geographically, I would like courses in the Southeastern USA, but will take
info on courses anywhere in the US.

John Giles
Senior Materials Engineer
Honeywell Space Systems
Clearwater, FL




From: Dale :      dshumake-at-welchlink.welch.jhu.edu
Date: Thu, 6 Mar 1997 13:38:49 -0500 (EST)
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe
dshumake-at-welchlink.welch.jhu.edu





From: PATRICK CLARK :      CLARK.PATRICK-at-EPAMAIL.EPA.GOV
Date: Thu, 06 Mar 1997 12:10:46 -0500
Subject: critical point dryers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The EM facility at EPA-Cincinnati is hopefully going to replace a 20 year
old critical point dryer in the near future. If anyone has any
recommendations, or horror stories, let us know before we spend your
tax dollars. You can reply to us directly at
clark.patrick-at-epamail.epa.gov Thanks




From: Bob_Citron-at-cc.chiron.com (Bob Citron)
Date: 3/6/97 8:34 AM
Subject: Al peak

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mark;

This happened to me once, and it turned out to be stray x-rays from the EDX
detector collimator sleeve. If your detector has one of these, slide it
off and paint the inside surfaces with carbon dag and try it again; it may
solve your problem.

Regards,

-Bob
**********************************
Bob Citron
Chiron Vision
555 W. Arrow Hwy
Claremont, CA 91711
ph: (909)399-1311
email: Bob_Citron-at-cc.chiron.com
**********************************

______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

When I perform an EDX analysis on uncoated fused silica, 10 20 or 30 KeV,
and the sample begins to charge, often I get a small aluminum peak, and it
doesn't go away as I move around the sample, it comes back. The amount of
aluminum in the quartz is typically less than 20 ppm, so I don't think that
the Al peak is coming from my sample.
Could it be that the Al peak is coming from my sample stage or walls of
the sample chamber, due to any effect caused by the charge that has developed
on the uncoated quartz?
Also, for EDX work, what are the negative effects that sample charging
has on an analysis? My work is done on an Amray 1600 with a Noran detector,
one that has two windows, Be and the ultra thin window.

Thanks, Mark Darus
Darus-at-cle.dnet.ge.com




From: bozzola-at-siu.edu (John J. Bozzola)
Date: Thu, 6 Mar 1997 16:17:53 -0600
Subject: Re: SEM - EDX references

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: bozzola-at-saluki-mail.fiber2.siu.edu
Message-Id: {v02130501af44f283ad5d-at-[131.230.97.73]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


Would you be so kind as to cite your literature references, so we can read
them.
They sound useful for a number of us doing EDX in a variable pressure SEM.
Thanks.

####################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Neckers Building, Room 146 - B Wing
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################






From: T. Robertson :      TROBERTS-at-eosin.path.uwa.edu.au
Date: Fri, 07 Mar 97 11:45:00 PST
Subject: Fixation of Dust Mite for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello out there in cyberspace
I am interested in examining dust mite at the TEM level. Has anyone got
any ideas for fixation for these beasts.

Terry Robertson

Dr Terry Robertson
Electron Microscopist
Department of Pathology
University of Western Australia
Nedlands 6009

phone 346 2935
Fax 346 2891
email troberts-at-eosin.path.uwa.edu.au




From: SONEJA A K :      soneja-at-giasbma.vsnl.net.in
Date: Fri, 7 Mar 1997 12:51:30 +0530 (IST)
Subject: Info on image analysis software/solutions for textiles/sperm analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello guys,

Looking for image analysis software/solutions for the following:

1} sperm analysis,cell motility etc.
2} textile analysis,rayon fibre analysis,broken fibre etc.

Could anyone point out some good manufacturers/sources with
address/email/fax wwith name of source.?//////
I would be grateful if someone could help me in this context.

Thanks,

Best regards,

Anish

*************************************************************************
For further details please contact:
Soneja A.K.
Director
METZER BIOMEDICAL & ELECTRONICS PVT.LTD.
327 Wadala Udyog Bhavan,Wadala,MUMBAI(BOMBAY )400 031.INDIA
Tel:91 22 4145057/4165650
Fax 91 22 4168757

Email:soneja-at-giasbma.vsnl.net.in
*************************************************************************





From: Ian Bache :      icb1000-at-cus.cam.ac.uk
Date: Fri, 07 Mar 1997 09:55:01 +0000
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There are a couple of reasons why windows can have a limited lifetime when
used in LV-SEM & ESEM.

Firstly, when the chamber is pumped to and from atmosphere any loose
particles in the chamber are disturbed, and can damage the window by
hitting it. This can be reduced by fitting some kind of loose cover (we
use a plastic bag!) over the x-ray probe whenever using the microscope for
non x-ray work.

Secondly, certain windows when used in lo-vac are susceptible to ice
formation on the surface and actually in the window.

Regarding probe-beam spreading, we, like Dr. Shah, are carrying out
monte-carlo simulations, in conjunction with x-ray measurements and will
hopefully be presenting some results at this years MSA meeting



Ian Bache
Cavendish Laboratory
Cambridge University
Madingley Road
Cambridge
ENGLAND
(+44)-1223-337229





From: chris gilpin :      CGILPIN-at-fs1.sem.man.ac.uk
Date: Fri, 7 Mar 1997 12:14:33 BST
Subject: Re: SEM - EDX references

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Would you be so kind as to cite your literature references, so we can read
} them.
} They sound useful for a number of us doing EDX in a variable pressure SEM.
} Thanks.
}
Sorry I should have included them on the original post!

Sigee, D.C. and Gilpin, C.J.
X-ray microanalysis with the environmental scanning electron
micrasocpe: Interpretation of data obtained under different
atmospheric conditions.
Scanning microscopy supplemnet 8 219-229 1994

X-ray microanalysis of wet biological specimens in the environmental
scanning electron microscope: 1 Reduction of specimen distance under
different atmospheric conditions.
Journal of Microscopy 179 (1) 22-28 1995



More work in progress as they say!

Chris





Chris Gilpin
Biological Sciences Electron Microscope Unit
G452 Stopford Building
Oxford Road
Manchester
M13 9PT
phone +44 161 275 5170
fax +44 161 275 5171




From: chris gilpin :      CGILPIN-at-fs1.sem.man.ac.uk
Date: Fri, 7 Mar 1997 12:14:33 BST
Subject: Re: SEM - EDX references

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Would you be so kind as to cite your literature references, so we can read
} them.
} They sound useful for a number of us doing EDX in a variable pressure SEM.
} Thanks.
}
Sorry I should have included them on the original post!

Sigee, D.C. and Gilpin, C.J.
X-ray microanalysis with the environmental scanning electron
micrasocpe: Interpretation of data obtained under different
atmospheric conditions.
Scanning microscopy supplemnet 8 219-229 1994

X-ray microanalysis of wet biological specimens in the environmental
scanning electron microscope: 1 Reduction of specimen distance under
different atmospheric conditions.
Journal of Microscopy 179 (1) 22-28 1995



More work in progress as they say!

Chris





Chris Gilpin
Biological Sciences Electron Microscope Unit
G452 Stopford Building
Oxford Road
Manchester
M13 9PT
phone +44 161 275 5170
fax +44 161 275 5171




From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Fri, 07 Mar 97 07:37:02 -0500
Subject: SEM Courses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

John Giles wrote:
===============================================

I have been tasked with bringing another technician into our SEM lab and
would like to get info on the courses which are available for a person who
is familiar with light microscopy but knows almost nothing about a SEM.

The desired course would be 5 days or less and be applicable to the level of
knowledge described above. We deal with strictly materials analysis, so
biological based courses are not applicable. The desired outcome would be
someone who has a rudimentary knowledge of SEM theory and capable of taking
basic SEM pictures and acquiring EDS spectra on non-challenging samples. We
will be doing some training of the new tech, but due to heavy workloads, we
can't devote the time to the training that we would normally do.
=================================================
We would give extremely high marks to the "Lehigh" courses which are given
at Lehigh University, Bethlehem, PA every year. You can get information
about their program from their website at the following:

http://www.Lehigh.EDU/~inmatsci/Microscourses.html

Chuck

===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: http://www.2spi.com
############################
==================================================




From: scott.wight-at-nist.gov (Scott Wight)
Date: Fri, 7 Mar 1997 10:22:25 -0500
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

We anticipated window problems from a detector left in the wet environment
24/7 and cycled for every sample change. We initially had a hard time
finding an EDS vendor to build us a detector that was fully retractable
behind a gate valve outside the 2020 chamber so that the detector was
protected not only from projectiles and cycles but from harsh environments
sometimes used in environmental scopes. We did eventually get such a
detector and had window problems initially but that seems to have been
solved with a slightly thicker window. My advice is that everyone should
let their EDS reps know that this type of detector is desirable and maybe
one day they will be widely available.

I also wonder if the commercially available window materials are water
tight. As we all know one of the EDS vendors uses a heater to remove ice
from its crystal. Helium leak tests are often quoted when asked about
water permeability but I would like to see water leak test data. If water
is geting into and through windows it could be responsible for lost windows
in wet systems.

I am looking forward to these monte carlo papers and lively discussions at
the M&M meeting in Clevland.

Scott Wight


} There are a couple of reasons why windows can have a limited lifetime when
} used in LV-SEM & ESEM.
}
} Firstly, when the chamber is pumped to and from atmosphere any loose
} particles in the chamber are disturbed, and can damage the window by
} hitting it. This can be reduced by fitting some kind of loose cover (we
} use a plastic bag!) over the x-ray probe whenever using the microscope for
} non x-ray work.
}
} Secondly, certain windows when used in lo-vac are susceptible to ice
} formation on the surface and actually in the window.
}
} Regarding probe-beam spreading, we, like Dr. Shah, are carrying out
} monte-carlo simulations, in conjunction with x-ray measurements and will
} hopefully be presenting some results at this years MSA meeting
}
}
}
} Ian Bache
} Cavendish Laboratory
} Cambridge University
} Madingley Road
} Cambridge
} ENGLAND
} (+44)-1223-337229

------------------------------------------------------------------
Scott Wight e-mail: SCOTT.WIGHT-at-NIST.GOV
NIST - Microanalysis Group W voice: 301-975-3949
Bld 222, Rm A113 | fax:301-216-1134/301-417-1321
Gaithersburg, MD 20899 \|/ disclaimer:
Any opinion expressed is my own and does not represent those of my employer.






From: Luc Nocente :      ln-at-noesisvision.com
Date: Fri, 07 Mar 1997 09:33:47 -0500
Subject: Re: Info on image analysis software/solutions for

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Try Visilog from Noesis Vision Inc, we have a turn key system available for
sperm analysis and or fibre analysis. You can reach us at
http://www.noesisvision.com




At 12:51 PM 3/7/97 +0530, SONEJA A K wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


----------------------------------------------------------------------------
---------------------
Luc Nocente Tel: 514 345 1400
Noesis Vision Inc. Fax: 514 345 1575
e-mail: ln-at-noesisvision.com
6800 Cote de Liesse, Suite 200
St-Laurent, PQ
H4T 2A7,Canada

Visit our new web site at http://www.noesisvision.com
----------------------------------------------------------------------------
---------------------




From: Rebecca Morden :      info-at-rms.org.uk
Date: Fri, 7 Mar 1997 15:54:56 +0000
Subject: Colloquium for Computers in Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Royal Microscopical Society
Computers in Microscopy ,September 1997, Cambridge
CALL FOR PAPERS

A Colloquium on various aspects of the application of computers to
microscopes and
microscopical techniques is being planned by the Royal Microscopical
Society, to be
held in the University Engineering Department, Trumpington Street,
Cambridge, on
Thursday 25th September 1997.

The colloquium is intended to cover recent progress made in the use of
computers for
the assessment, interpretation, restoration and enhancement of
microscopical images,
including, but not restricted to applications in optical and all kinds
of electron
microscopy. Of particular interest will be contributions concerning new
approaches
and techniques including 3D measurements and profiling, wavelets and
fractals.
It is hoped that papers will cover some of the comparatively new fields
of application,
including multi-channel imaging, instrumental control and remote
microscopy. It is
proposed also to organise a small exhibition of products and materials
by local
organisations involved in this area, as well as a visit to laboratories
in the University
engaged in this kind of work.

The colloquium takes place immediately after a three-day course, also
entitled
'Computers in Microscopy', and it is intended that the two events be
complementary.
( Please note that it is not necessary to attend the course in order to
register
for the colloquium.)

If you would like further information about the Colloquium and/or the
Course please
contact Rebecca Morden either by email: info-at-rms.org.uk,
by telephone: +44 (0)1865 248768 or by fax: +44 (0)1865 791237.

Prospective contributors are invited to submit a synopsis of
approximately 150-200
words, before 30 April 1997, to Dr D M Holburn, University Engineering
Department,
Trumpington Street, Cambridge CB2 1PZ, U.K. Enquiries may be made by
electronic
mail, to dmh-at-eng.cam.ac.uk, by facsimile to +44 1223 332662, or by
telephone to
+44 1223 332775.

*******************************************************************************
Rebecca Morden
Course Organiser
Royal Microscopical Society, 37/38 St Clements, Oxford, OX4 1AJ, UK.
Tel (0)1865 248768, Fax (0)1865 791237, email info-at-rms.org.uk
*******************************************************************************




From: Zhiyu Wang :      wangz-at-pulsar.cs.wku.edu
Date: Fri, 07 Mar 1997 10:57:16 -0600
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello Ian and all:

Thank you very much for providing information and experience in window
breackage problem.
I got windows replace service with the price of $6500 (does not
including freight charges).
We do not know how to handle this problem if the window does not stand
in a reasonable lifetime
or unlucky, got several window breackages. It must be a big meney gone!
Is it possible to find a cheaper service? Any suggestion?

Again, thank you very much for help.

Zhiyu




From: Giles John E Jr :      giles_john_e_jr-at-space.honeywell.com
Date: Fri, 7 Mar 97 11:11:42 CST
Subject: SEM Courses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Everyone,

I have been tasked with bringing another technician into our SEM lab and would
like to get info on the courses which are available for a person who is
familiar with light microscopy but knows almost nothing about a SEM.

The desired course would be 5 days or less and be applicable to the level of
knowledge described above. We deal with strictly materials analysis, so
biological based courses are not applicable. The desired outcome would be
someone who has a rudimentary knowledge of SEM theory and capable of taking
basic SEM pictures and acquiring EDS spectra on non-challenging samples. We
will be doing some training of the new tech, but due to heavy workloads, we
can't devote the time to the training that we would normally do.

Geographically, I would like courses in the Southeastern USA, but will take
info on courses anywhere in the US.

John Giles
Senior Materials Engineer
Honeywell Space Systems
Clearwater, FL




From: kbart-at-hamilton.edu (Ken Bart)
Date: Fri, 7 Mar 1997 12:51:45 -0500
Subject: Re: SEM Courses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} John Giles wrote:-----------------------------------------------.
}
} Hello Everyone,
}
} I have been tasked with bringing another technician into our SEM lab and would
} like to get info on the courses which are available for a person who is
} familiar with light microscopy but knows almost nothing about a SEM.
}
} The desired course would be 5 days or less and be applicable to the level of
} knowledge described above. We deal with strictly materials analysis, so
} biological based courses are not applicable. The desired outcome would be
} someone who has a rudimentary knowledge of SEM theory and capable of taking
} basic SEM pictures and acquiring EDS spectra on non-challenging samples. We
} will be doing some training of the new tech, but due to heavy workloads, we
} can't devote the time to the training that we would normally do.
}
} Geographically, I would like courses in the Southeastern USA, but will take
} info on courses anywhere in the US.
}
} John Giles
} Senior Materials Engineer
} Honeywell Space Systems
} Clearwater, FL


John:

A number of us run an SEM course at the University of Maryland
called 'Practicle Aspects of SEM. This course is a 4 1/2 day course
covering the basic theory and practice of SEM (including basic x-ray
microanalysis) with direct hands-on experience for all students. In
addition, students are encouraged to bring samples of their own. This year
the courses are set for May 19-23 and May 26-30. If you would like more
information please feel free to contact either me ( 315-859-4715) or Tim
Maugel (301-405-6898)

Ken Bart

Kenneth M. Bart
Director, Electron Microscopy Facility
Hamilton College
Clinton, New York 13323 USA
kbart-at-hamilton.edu
(315) 859-4715






From: Weiland, Hasso :      Hasso.Weiland-at-alcoa.com
Date: Fri, 7 Mar 1997 13:54:13 -0500
Subject: Vibration Isolation Pad

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

We are in need of a vibration isolation pad for an Auger spectrometer,
which has the dimensions of a standard SEM. Can anybody recommend a
manufacturer or can recommend homemade solutions?




From: Weiland, Hasso :      Hasso.Weiland-at-alcoa.com
Date: Fri, 7 Mar 1997 14:37:40 -0500
Subject: FS: WDS system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

We have a WDS system for sale. The system consists of the computer,
software, spectrometer controls and stage controls, NO DETECTORS or
SPECTROMETERS are included.

It is a Kevex Delta 4 / Sesame system (type 8006) with a 5724-004
console, containing 4 spectrometers controls, 3 stage motors, and 4
spectrometer pre-amps and ratemeters, beam meter. A Tectronix plotter
and a matrix printer a part of the package.

The system was configured to run on a JEOL JXA 8600 superprobe.

Price: $ 5000, you pay freight.


Hasso Weiland
Alcoa Technical Center
Alcoa Center, PA 15069

412 337-3133 (phone)
412 337-2044 (Fax)
hasso.weiland-at-alcoa.com







From: Andrew Buechele :      andrew-at-rsrch.vsl.cua.edu
Date: Fri, 7 Mar 1997 14:43:09 EST
Subject: Re: Printheads for Color Stylewriter Pro

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I hope this doesn't seem out of palce on this list, but it is
equipment I need to print spectra and images from ny TEM work. Has
anyone had to replace the printhead on an Apple Color StyleWriter Pro
and gotten a better price than the going Apple rate of $250! (new
printer time?) Alternately, has anyone got a unit that was junked for
other reasons that might have a usable printhead that they would be
willing to sell/donate? The one that I use in my electron microscope
lab had a malfunction which forced ink on top of the printhead and
corroded out the contacts. Thanks in advance for any help/advice
you can offer.

Sincerely,
Andy
Andy Buechele
The Catholic University of America
409 Hannan Hall
Washington, D.C. 20064
(202) 319-4995 FAX: (202) 319-4469




From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Fri, 7 Mar 1997 13:10:46 -0800 (PST)
Subject: CPD Help!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello All,


Y'all are an encyclopedia of knowledge, and I want to check out a
volume... I have just recently been learning SEM and all the associated
procedures that precede going on the scope. I have had to learn this all
on my own, which is why I'm having this problem.
When I do a cpd on a sample they always come out looking wrinkled.
Am I going wrong in the cpd or is it some other step. I've heard that some
people do a few purges of the system and some do a lot. I've just purged
the ETOH until I see CO2 snowballs & then let it sit for 15 min. & then
repeat the cycle 2 more times. Is this wrong?
Can someone out there give me some pointers? I'm getting pretty
tired of my stuff looking like raisins all the time. = (


Thanks in advance!

Paula = )

Paula Ssicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu






From: moxtek-at-MOXTEK.WIN.NET (Clark Turner)
Date: Fri, 07 Mar 1997 14:32:19
Subject: Re: SEM - EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Microscopy List,

This seems to be a question of general interest. As the
manufacturer of the majority of ultrathin windows, MOXTEK can
offer the following suggestions to improve window lifetimes.

1) Slow down the chamber venting. The most common field failure
is what we call "bullet holes." These occur when particulate in
the chamber is "swirled up" during venting and impacts the surface
of the window.

2) Retract the detector before you vent, if you have this
capability. (See #1 for the reason.) I saw an earlier suggestion
to cover the detector with a plastic bag when it is not in use to
minimize the exposure to this flying particulate. Good suggestion!

3) Never pump down your specimen chamber with a warm x-ray
detector attached. Make sure you cool down the detector before
pumpdown. Most of the detector systems have a molecular sieve or
other getter material in the dewar vacuum to trap any residual
gases. When you warm up the detector this material can outgas the
trapped gases. If you then pump down the specimen chamber you
reverse pressure the window. These windows are not designed to
tolerate any back pressure.

4) Never overheat the window by exposure to a hot stage. Your
instrument manufacturer should be able to recommend how to operate
the system to minimize heat transfer to the detector window.

5) Consult with your EDX manufacturer for specific instructions
for your system. Every detector/microscope combination is unique,
and they should have some experience that will help you.

There are various thicknesses of windows available, but the above
suggestions apply regardless of the thickness. If you continue to
have failures with the thinnest windows, a thicker option is
probably available through your EDX manufacturer.

(Please note that the Kevex B windows are unique to Kevex, so you
should probably talk to them about the problem you had. With a
silicon support grid like the MOXTEK ultrathin windows there is no
problem in frequent pressure changes breaking the window.)

Hope this helps.

D. Clark Turner
Director
MOXTEK, Inc.
452 West 1260 North
Orem, Utah 84057
phone (801) 225-0930
fax (801) 221-1121
email moxtek-at-moxtek.win.net

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Roar Kilaas :      Roar_Kilaas-at-macmail.lbl.gov
Date: 7 Mar 1997 17:32:13 -0700
Subject: Job Posting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

REGARDING Job Posting

POSITION OPEN: Postdoctoral Scholar or Research Associate

Area:
Materials Science/Analytical Electron Microscopy/Lorentz Imaging

Qualification:
A PhD in Materials Science/Physics or related area. Very strong hands-on
experience in various TEM techniques and their application to microstructural
characterization at high resolution is required. Experience in Lorentz
Imaging and TEM specimen preparation is desirable. The suitable candidate
will be comfortable using the wide range of TEM instrumentation available at
the NCEM.
This post-doctoral fellow will spend a major fraction of the time working on
the microstructural characterization of plastically deformed samples. The
goal is to explore the correlation between the local physical/chemical
microstructure in plastically deformed regions with local changes in magnetic
structure(measured independently by SQUID imaging techniques). The
remaining time will be devoted to the development of Lorentz Imaging
techniques on our new Philips 200KV/FEG instrument using a variety of thin
film samples.

The position is open immediately for at least one year, renewable
upon mutual agreement for longer period, provided funds are available. Salary
and benefits will commensurate with experience and skills.

For more information on the scientific/technical aspects of the
position please contact :

Kannan Krishnan
72-209, NCEM
Lawrence Berkeley National Laboratory
Berkeley, CA 94720

Krishnan-at-LBL.GOV
Tel: 510-486-4614
Fax: 510-486-5888







From: Woody.N.White-at-mcdermott.com
Date: Fri, 7 Mar 1997 8:10:00 -0600
Subject: Re: SEM Courses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lehigh University, Bethlehem, PA has beginner as well as advanced
SEM/EDS (+other) courses. These run for a week, and are well done.
Sorry, I tossed my mailing about the course so I am missing the
details. Believe it is scheduled for June. I am sure someone else
will jump in with the details.

There are other (similar) courses available, but I have not
experienced those....

Woody White




From: Woody.N.White-at-mcdermott.com
Date: 3/6/97 11:03 AM
Subject: Carbon analysis quantification using EDX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Quantitative (even standardized) EDS analysis for carbon in Nb is
difficult to impossible, no matter what the EDS manufacurers may
imply. The same applies for carbon in Zr. Several things.....
1) Very light element (soft x-rays) in a high Z matrix.
2) Nb is a strong absorber of carbon x-rays.
3) Since the matrix carbon signal is weak, any surface carbon
(contamination) can be a major contributor to the carbon peak.
Together, these make for very large absorption correction coefficients
and magnify surface contamination contributions.
Translate that: error.

To maximize your chances....

Don't rely on standardless (semi) quant.

A clean, high vacuum is required to minimize surface (carbon) during
analysis.

Specimen surface preparation is crucial. A flat, well polished, and
very clean specimen surface must be achieved.

Do not coat the specimen (NbC is sufficiently conductive anyway).

Minimize analysis volume and depth. - (A) Use as low a beam voltage as
practical. (B) High tilt angles will skew the volume to near surface.
Caveats: (A) Nb, L-line (family) response less predictable than Ka
line. I have run across some quant software which won't use L-lines-
at least for stdless quant. (B) Less than perfect quant compensation
for high tilt may add some error...(especially for stdless).

For stdless quant, I have seen demos produce errors } 300 percent.

Good luck! Let me know how things work out.

Woody White


______________________________ Reply Separator
_________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I am trying to do trace Carbon analysis in Nb material. The Be window of my
EDX can be opened in order to detect C, N, O, but the quantification seems
impossible. Anybody has similar experience?

Any response is appreciated!

Best regards.

Tong




From: Mary Mager :      mager-at-unixg.ubc.ca
Date: Fri, 07 Mar 1997 20:13:20 -0800
Subject: Re: Vibration Isolation Pad

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {2.2.32.19970308041320.006ac864-at-pop.unixg.ubc.ca}
X-Sender: mager-at-pop.unixg.ubc.ca
X-Mailer: Windows Eudora Pro Version 2.2 (32)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Dear Hasso,
I have had some success reducing the vibration evident on my SEM by making
four pads from Sorbathane. I used a fairly stiff number of Sorbathane about
3/8 inch thick and cut four pads about four inches square. The shop made me
eight steel plates to go on either side of the Sorbathane and I put these
under the corners of the column section. The vibration on my SEM (from being
on the fourth floor) went down about half. It may not be enough for Auger
but it is inexpensive. TMC, who advertises in most JMSA journals, has a full
set of platforms. Phone:508-532-6330, fax:508-531-8682. I have no experience
with their service.
You wrote:

} We are in need of a vibration isolation pad for an Auger spectrometer,
} which has the dimensions of a standard SEM. Can anybody recommend a
} manufacturer or can recommend homemade solutions?
}
Best of luck,
Mary
Mary Mager
Electron Microscopist
Metals and Materials Eng., UBC
6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel:604-822-5648, fax:604-822-3619
e-mail: mager-at-unixg.ubc.ca





From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Sat, 8 Mar 1997 07:50:58 +0000
Subject: Re: CPD Help!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Y'all are an encyclopedia of knowledge, and I want to check out a
} volume... I have just recently been learning SEM and all the associated
} procedures that precede going on the scope. I have had to learn this all
} on my own, which is why I'm having this problem.
} When I do a cpd on a sample they always come out looking wrinkled.
} Am I going wrong in the cpd or is it some other step. I've heard that some
} people do a few purges of the system and some do a lot. I've just purged
} the ETOH until I see CO2 snowballs & then let it sit for 15 min. & then
} repeat the cycle 2 more times. Is this wrong?
} Can someone out there give me some pointers? I'm getting pretty
} tired of my stuff looking like raisins all the time. = (
}
}
} Thanks in advance!
}
} Paula = )
}
} Paula Ssicurello
} UC Berkeley
} Electron Microscope Lab
} psic-at-uclink4.berkeley.edu

The main problem particularly newcomers to CPD have is trying to do
everything too quickly. I'm no expert at CPD having only done it a few
times, but I have been told:

a) when flushing, it need to be done SLOWLY - if the CO2 boils (or even
simmers gently), then the specimen will be wrecked.

b) It is important to leave your specimen in the liquid CO2 for long enough
for proper infiltration, and that this needs doing at least two or three
times to make sure all the acetone/alcohol is displaced - even for small
specimens (1mm), this can take 3-4 hours. Larger specimens might require 24
hrs.

c) At the end, when releasing the CO2 gas, again it must be done slowly -
15-20 mins. Sudden pressure releases will blow the specimen apart.

Regards,

---------------------------------------------------------------
Dr L. P. Stoter Technical Editor, MICROSCOPY & ANALYSIS
Technesis
17, Rocks Park Road email: LPS-at-teknesis.demon.co.uk
Uckfield, E. Sussex Phone: +44 (0)1825 766911
TN22 2AT Fax: +44 (0)1825 766911
United Kingdom
---------------------------------------------------------------






From: wise-at-vaxa.cis.uwosh.edu
Date: Sat, 08 Mar 1997 12:35:22 +0000
Subject: Supplies and teaching

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Thanks to all who responded to my question about teaching supplies. I
received about 20 responses. They fell into three basic types 1) give
each student an allotment at the beginning of the semester and require them
to buy more if they use it up; 2) charge each student a flat fee and let
them use at will; 3) require them to but their own supplies at a local
camera store. Some labs separated EM film (which can be hard to purchase
locally) from paper (which is stocked locally).

I'm leaning towards using options #1 or #3.

Thanks again.

Bob


Robert R. Wise
Plant Physiologist and Director, UWO Electron Microscope Facility
Department of Biology
University of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404 tel
(414) 424-1101 fax
wise-at-uwosh.edu






From: Edmund Glaser :      eglaser-at-umabnet.ab.umd.edu
Date: Sun, 9 Mar 1997 01:36:19 -0500 (EST)
Subject: Re: Morphometric Systems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

You should look into Neurolucida by MicroBrightField. It is a 3D
morphometric analysis system. MBF also is also active in stereology
analysis and was an exhibitor at the Soc. Neurosci meeting. You can
get more info from their web site at www.microbrightfield.com/microb/ or
via email at info-at-microbrightfield.com.

On Thu, 27 Feb 1997, Cheri Owen
wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
}
} A friend of a friend is looking for a morphometric system with Scope,
} stylus, PC etc. They saw some at the recent Neurosci meetings, but didn't
} keep the brochures. Any info would be greatly appreciated.
}
} Thanks
}
} Cheri Owen
} Detroit Neurotrauma Institute
} Wayne State University
} Detroit, MI
}
}
}

Edmund Glaser, D. Eng.
Dept. Physiol.
Univ. Md. School. Med.
Baltimore, MD 21201 USA
Ph: (410) 706-5041
Fax: (410) 706-8341





From: D.Wild-at-mirinz.org.nz
Date: Mon, 10 Mar 1997 09:15 +1200
Subject: LM photomicrography

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

We are trying to photograph bacteria using phase contrast optics. We
are having trouble getting a thin enough layer to obtain 'fully
focused" micrographs at 100x under oil. We are using a fixed
suspension mixed with "aquatex" mounting medium.
Any suggestions?
Thanks David Wild




From: Scott D. Walck WL/MLBT :      walcksd-at-ml.wpafb.af.mil
Date: Sun, 9 Mar 97 20:14:14 -0500
Subject: Attention MRS Spring '97 Symposium Z authors-TEM Sample Prep.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am sorry that this is a little late. I will be trying to get this message
out to the individual authors who will be presenting who do not receive it
through the Listserver. If you are the principal author or a coauthor on a
paper being presented at Symposium Z at the Spring '97 MRS meeting, please let
me know that you received this anouncement ASAP.

The papers that you are submitting to the "Workshop on TEM Specimen
Preparation-IV", Symposium Z should be tutorial in nature. They should be
written in a "how I did it and here's how you can do it" mode with all the
tidbits and tricks-of-the-trade that make your technique work. Follow the
authors guidline for the paper for the MRS proceedings except for length. Use
as many pages as you need to tell your story. Use as many pictures and
diagrams that you need to get the information across. If you have any
questions, contact me or Ron Anderson.

- -Scott Walck

Ron Anderson, Chair
IBM, ZIP E-70
East Fishkill Facility
Hopewell Jct., NY 12533
*(914)892-2225 (-2003 FAX)
ron-anderson-at-vnet.ibm.com

Scott D. Walck, Co-Chair
Materials Directorate, Wright Laboratory
Wright Patterson AFB, OH 45433-7750
*(513)255-5791 (-2176 FAX)
walcksd-at-ml.wpafb.af.mil




From: MEV H VAN DER MERWE :      HVDM-at-op1.up.ac.za
Date: Mon, 10 Mar 1997 09:53:00
Subject: Immuno-Bed staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Can anyone please tell me the name of staining books in the field of
Immuno-bed tissues as well as the companies from which I can order it.
We are interested in
- Histology and Pathology
- Immunofluorescence
- Immunohistochemistry
- Histofluorescence
- Enzymehistochemistry

Thanks
Hildagonda van der Merwe

University of Pretoria
Faculty of Veterinary Science
Dept. of Pathology
Onderstepoort
0110

tel: 012-5298176
fax: 012-5298303
e-mail:Hvdm-at-op1.up.ac.za




From: greg :      greg-at-umic.sunysb.edu
Date: Mon, 10 Mar 1997 09:55:27 +0000
Subject: Re: CPD Help!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello All,
}
}
} Y'all are an encyclopedia of knowledge, and I want to check out a
} volume... I have just recently been learning SEM and all the associated
} procedures that precede going on the scope. I have had to learn this all
} on my own, which is why I'm having this problem.
} When I do a cpd on a sample they always come out looking wrinkled.
} Am I going wrong in the cpd or is it some other step. I've heard that some
} people do a few purges of the system and some do a lot. I've just purged
} the ETOH until I see CO2 snowballs & then let it sit for 15 min. & then
} repeat the cycle 2 more times. Is this wrong?
} Can someone out there give me some pointers? I'm getting pretty
} tired of my stuff looking like raisins all the time. = (
}
}
} Thanks in advance!
}
} Paula = )
}
} Paula Ssicurello
} UC Berkeley
} Electron Microscope Lab
} psic-at-uclink4.berkeley.edu

Paula,
It sounds like your samples are still "wet". Make sure
that your are dehydrating your samples long enough. The
ETOH or acetone must be 100% pure. The time you are
alloting in the CPD is too short. This will also leave
your sample wet with acetone or ETOH. When you drain the
ETOH do it slowly and place a dry Kimwipe close and infront
of the vent. When it no longer gets wet after a few purges
the ETOH is gone. Then Let it sit for 30 min. more.
Bring the temp. of the chamber up to about 37C slowly.
Vent the CO2 over 10 min until the pressure is zero.
The above is all sample dependent and must be worked out
by trial and error. Plant material will take longer than a
cell cultuer on a cover slip.

Greg Rudomen
University Microscopy Imaging Center
S.U.N.Y. Stony Brook
Greg-at-umic.sunysb.edu





From: David Strecker :      strecker-at-bright.net
Date: Mon, 10 Mar 1997 10:52:17 -0500
Subject: 5 days left for abstract submissions!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

This is to remind everyone that extended abstracts for Microscopy &
Microanalysis '97 need to be subimtted by MARCH 15! That is this
Saturday! Papers received after that date will not be accepted.

If you have misplaced or have not received a bulletin or need more
information, you can contact

Microscopy & Microanalysis '97
4 Barlows Landing Rd., Suite 8
Pocasset, MA 02559
phone: 508-563-1155
Toll-free: 800-538-3672
FAX: 508-563-1211

Email: BusinessOffice-at-MSA.Microscopy.com or
WWW: http://www.MSA.Microscopy.com

Meeting information is also available at
http://www.bright.net/~strecker/msno/mm97.html


Thanks,
-Dave Strecker




From: Scott D. Walck WL/MLBT :      walcksd-at-ml.wpafb.af.mil
Date: Mon, 10 Mar 97 11:02:52 -0500
Subject: Author Guidlines for MRS '97 Symposium Z__TEM Sample Prep

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Poster instructions can be found at the MRS web site, www.mrs.org

You should have recieved author instructions. Please contact MRS if you have
not. The Materials Research Society, 9800 McKnight Road, Pittsburgh, PA
15237-6006
Phone: 412/367-3004, Fax: 412/367-4373, Email: info-at-mrs.org

If you haven't, this should get you started.

For camera-ready text, on an 8-1/2 x 11 inch paper, the margins are
left: 1 inches
right 1 inches
top: 1/2 inch
bottom: 1 inch
Page number at 1/2 inch from bottom of page.

That gives 6.5 inch wide x 10 inch box to put your text, single spaced, 12
point font Times Roman. Use the AIP reference style or the MRS refernce
style. The manuscripts will be reduced by 26%. Scale markers on all
micrographs. Check out other proceedings for style if you haven't received
your instructions.

Bring the original manuscript and (3) photocopies to the meeting. You will
need to have the MRS copyright form filled out. Check the bulletin boards on
where to take your paper. Ron or myself will be collecting them from you and
will post an announcement.

- -Scott


*****************************************************************************
Scott D. Walck, Ph.D. *
Materials Directorate Tel: (937) 255-5791 *
2941 P St Ste 1 Fax: (937) 255-2176 *
WL/MLBT, BLDG 654 Home: (937) 427-1093 *
Wright Patterson AFB, OH 45433-7750 *
*
EMAIL: *
Work: walcksd-at-ml.wpafb.af.mil Home: SKAB-Walck-at-worldnet.att.net *
*****************************************************************************





From: Rebecca Morden :      info-at-rms.org.uk
Date: Mon, 10 Mar 1997 16:20:06 +0000
Subject: Computers in Microscopy Course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The Royal Microscopical Society
Computers in Microscopy ,September 1997, Cambridge
CALL FOR PAPERS

A Colloquium on various aspects of the application of computers to
microscopes and
microscopical techniques is being planned by the Royal Microscopical
Society, to be
held in the University Engineering Department, Trumpington Street,
Cambridge, on
Thursday 25th September 1997.

The colloquium is intended to cover recent progress made in the use of
computers for
the assessment, interpretation, restoration and enhancement of
microscopical images,
including, but not restricted to applications in optical and all kinds
of electron
microscopy. Of particular interest will be contributions concerning new
approaches
and techniques including 3D measurements and profiling, wavelets and
fractals.
It is hoped that papers will cover some of the comparatively new fields
of application,
including multi-channel imaging, instrumental control and remote
microscopy. It is
proposed also to organise a small exhibition of products and materials
by local
organisations involved in this area, as well as a visit to laboratories
in the University
engaged in this kind of work.

The colloquium takes place immediately after a three-day course, also
entitled
'Computers in Microscopy', and it is intended that the two events be
complementary.
( Please note that it is not necessary to attend the course in order to
register
for the colloquium.)

If you would like further information about the Colloquium and/or the
Course please
contact Rebecca Morden either by email: info-at-rms.org.uk,
by telephone: +44 (0)1865 248768 or by fax: +44 (0)1865 791237.

Prospective contributors are invited to submit a synopsis of
approximately 150-200
words, before 30 April 1997, to Dr D M Holburn, University Engineering
Department,
Trumpington Street, Cambridge CB2 1PZ, U.K. Enquiries may be made by
electronic
mail, to dmh-at-eng.cam.ac.uk, by facsimile to +44 1223 332662, or by
telephone to
+44 1223 332775.


*******************************************************************************
Rebecca Morden
Course Organiser
Royal Microscopical Society, 37/38 St Clements, Oxford, OX4 1AJ, UK.
Tel (0)1865 248768, Fax (0)1865 791237, email info-at-rms.org.uk
*******************************************************************************




From: wwiggins-at-mail.carolinas.org
Date: 3/10/97
Subject: TEM Job Post - 10 Mar 1997

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-------------------------------------
Name: Winston Wiggins
E-mail: wwiggins-at-carolinas.org

Job Number: 73252-RM

Location : Carolinas HealthCare System
Charlotte, North Carolina, USA

Shift: Full-time, M-F, 8a-4:30p.

Essentials: Fix, dehydrate, and embed biological tissue specimen for transmission e.m.
Thick and thin sectioning, Basic operation of Philips CM10 electron microscope.
Develop and print photomicrographs. Maintain logs and records of work,

Requirements: BA/BS degree in related area. Minimum 1.5 year relatively recent experience in
electron microscopy, preferably in a work setting. Good "people skills."

Compensation: Complete benefits package, relocation expenses.

Contact: Roxy McKinney
Carolinas HealthCare System
Human Resources
P.O. Box 32861
Charlotte, N.C. 28232-2861
USA

Fax:1 704 355-1728

NOTA BENE: Position is guaranteed for one year but with mutual consent and expected
continued funding will become permanent afterward.






From: Andrew Buechele :      andrew-at-rsrch.vsl.cua.edu
Date: Mon, 10 Mar 1997 13:02:08 EST
Subject: Re: Vibration Isolation Pad

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You Wrote:

} We are in need of a vibration isolation pad for an Auger spectrometer,
} which has the dimensions of a standard SEM. Can anybody recommend a
} manufacturer or can recommend homemade solutions?

There are some low cost pneumatic mounts available from Barry
Controls in a wide range of load handling capacities. The trade name
is Stabl-Levl. I used these to successfully isolate a heavy FTIR unit
that was sensitive to vibration. Unless you can attach them directly
to your instrument, however, you will still need a stiff plate to set
it on with the pneumatic mounts between the plate and the floor. The
address and phone # I have for Barry Controls is:
700 Pleasant Street
Watertown, Mass. 02172
617-923-1150
Haven't ordered from them in a few years, so you might need to check
it if you can't connect. Disclaimer: I have no interest in Barry
Controls other than being a satisfied customer.

Andy Buechele
The Catholic University of America
409 Hannan Hall
Washington, D.C. 20064
(202) 319-4995 FAX: (202) 319-4469




From: Walter A. Mannheimer :      wamann-at-metalmat.ufrj.br
Date: Mon, 10 Mar 1997 15:17:09 EST3EDT
Subject: vibration isolation pad

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Our Auger with scanning tunnel microscope (with console)hangs from
ceiling beams on three springs
if you need details you may contact Prof. Achete
(achete-at-metalmat.ufrj.br) in this department
Best regards
Prof.Walter A.Mannheimer
Metallurgy and Materials Engineering
Federal University of Rio de Janeiro
POBox 68505 21945 Rio de Janeiro, Brazil
Voice +5521 280-7443 (Dept.office) +5521 590-0579 (direct)
Fax +5521 290-6626 Email: wamann-at-metalmat.ufrj.br




From: John F. Mansfield :      jfmjfm-at-umich.edu
Date: Mon, 10 Mar 1997 16:53:39 -0500
Subject: Mounting samples for hot stage work

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi there everyone.

Can anyone tell me the best way to mount a sample for hot stage work. We
have a cross section of a silicon nitride sample that we want to examine at
elevated temperatures (up to about 1000 deg C, if our hot stage will get us
there!) The sample is a cross section made by tripod polishing and it must
be mounted on a Mo or Ta washer for insertion into the scope, how are we
going to "glue" it to the ring so that the "glue" wont decay and the sample
fall off or move about at high temp? What we need is a very low vapor
pressure adhesive that can withstand high temperatures!

Any help appreciated. Thanks.

Jfm.

John Mansfield
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (313) 936-3352 FAX (313) 936-3352
Cellular Phone: (313) 715-2510
(Leaving a phone message at 936-3352 is preferable to 715-2510)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html






From: Raymond F Egerton :      egerton-at-phys.ualberta.ca
Date: Mon, 10 Mar 1997 16:26:12 -0700 (MST)
Subject: MSC conference abstracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Due to a projected streamlining of the review and production process,
abstracts for the June meeting of the Microscopy Society of Canada can be
accepted up to 2 weeks later than previously advertised, i.e. they must
be received in Edmonton by 1 April.

The deadline for conference pre-registration is 1 May 1997, as generally
advertised (but NOT 1 March as stated on the registration form!)

Ray Egerton, Physics Dept, University of Alberta, Edmonton, Canada T6G 2J1
Phone: 403-492-5095, FAX: 403-492-0714, e-mail: egerton-at-phys.ualberta.ca
------------------------------------------------------------------------






From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Mon, 10 Mar 1997 16:11:10 -0800 (PST)
Subject: Glove Results?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello All!


Afew weeks ago, there was someone who was doing a survey of the
best types of gloves to use in an EM lab. Of course I didn't save the
information passed on, and of course now we're interested in those results
here in our lab.
If anyone has those results, please let me know. Also, we work
with a lot of the low-temperature resins (Lowicryls, LR Gold), if anyone
knows of the best gloves to work with that stuff let me know that too.


Thanks oodles!

Paula = )

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu






From: wong-at-msg.ucsf.edu (Mei Lie Wong)
Date: Mon, 10 Mar 1997 17:21:17 -0700
Subject: Balzer's freeze-fracture

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like tofind someone in the S.F. Bay area that has a freeze-fracture
machine that would be available for doing samples. I have a colleague in
the area who is looking for a facility where she might be able to do some
samples.
We would be very grateful for any leads. Thanks in advance. ML

Mei Lie Wong
Department of Biochemistry
HHMI-UCSF
Ph. 415-476-4441 Fax 415-476-1902
email wong-at-msg.ucsf.edu






From: mgb-at-nucleus.ansto.gov.au (Mark Blackford)
Date: Tue, 11 Mar 1997 16:14:29 +1200
Subject: Reconditioning sputter ion pump elements

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am about to replace the sputter ion pump elements in our two JEOL TEM's.
JEOL used to be able to recondition the elements used in their 2000FX
microscopes but no longer provide this service. Can anyone suggest another
company that does this sort of work? How long can I expect a reconditioned
unit to last compared with a new one and what sort of cost would be
reasonable given that JEOL are charging $2.5k and $4.5k (Australian
dollars) for new units? Thanks in advance for any advice you can give.

Mark Blackford
TEM Group
Materials Division, ANSTO
PMB 1,
Menai, N.S.W.
Australia
2234
Phone 61 2 9717 3027
Fax 61 2 9543 7179

Disclaimer:
The views expressed in this E-mail message do not necessarily represent the
official views of ANSTO from which this message was conveyed.






From: Maria do Carmo Goncalves :      maria-at-iqm.unicamp.br
Date: Tue, 11 Mar 1997 11:28:44 -0300 (EST)
Subject: staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Please send me information on how to stain enzymes for electron microscopy
(either SEM or TEM)

Dr. S. P. Nunes




From: Gregory.Argentieri-at-sandoz.com (by way of Nestor J. Zaluzec)
Date: Tue, 11 Mar 1997 07:45:33 -0500
Subject: Stereology and Morphometry Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of any Stereology and Morphometry software that is
made for DOS/Windows OS?

Gregory Argentieri
Novartis Pharmaceutical Corp.
East Hanover, NJ
Gregory.Argentieri-at-Sandoz.com






From: Gregory.Argentieri-at-sandoz.com (by way of Nestor J. Zaluzec)
Date: Tue, 11 Mar 1997 07:45:33 -0500
Subject: Stereology and Morphometry Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of any Stereology and Morphometry software that is
made for DOS/Windows OS?

Gregory Argentieri
Novartis Pharmaceutical Corp.
East Hanover, NJ
Gregory.Argentieri-at-Sandoz.com






From: lizard-at-okway.okstate.edu (Ginger Baker)
Date: Tue, 11 Mar 1997 08:59:38 -0600
Subject: Oklahoma Microscopy Society Spring Workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The Oklahoma Microscopy Society (OMS) will be hosting its 19th Annual
Spring Workshop on Friday, April 4, 1997.

Topic: FORENSIC MICROSCOPY



MORNING SESSION: 129 George Cross Hall, University of Oklahoma in
Norman, OK.

8 - 9 am: Registration

9 - 9:05: Greetings

9:05 - 9:50: Guest speaker: Mark Betts (from Oxford Instruments) will
speak on "Microanalysis in the Forensic Laboratory"

9:50 - 10: Break

10 - 12: Guest speaker: Keith Ferrell (from the Oklahoma State Bureau
of Investigation (OSBI)) will speak on "Overview of Forensic
Microscopy"

12 - 1:30pm: Lunch/Executive Meeting/Open Meeting



AFTERNOON SESSION: OSBI Central Laboratory, 2132 N.E. 36th St. in
Oklahoma City

1:30 - 2: Travel to the OSBI

2 - on: Tours of the OSBI Facility



REGISTRATION (includes lunch):

You can download a copy of the registration form by visiting the OMS
website at http://www.ou.edu/research/electron/oms/ maintained by
Dr. Scott Russell, Nobel Electron Microscopy Lab at the University of
Oklahoma.

OMS members: the Newsletter was sent out Monday, March 10, 1997 so you
should be receiving it this week. A registration form/envelope has
been included in the newsletter. Please return it as soon as possible
so as to estimate numbers for lunch and parking permits.


Student and Professional Members: $5
Student Non-Members: $10
Professional Non-Members: $15

Pre-register until April 2, 1997 with Phoebe Doss
OMS President Elect
Dept. of Plant Pathology
110 NRC
Oklahoma State University
Stillwater, OK 74078
(405) 744-7995
Email: pjdoss-at-okway.okstate.edu

On site registration available.




From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Tue, 11 Mar 1997 11:40:21 -0800
Subject: re:staining enzymes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Dr. Nunes:

Your question, "how to stain enzymes for EM" is so broad as to be
unanswerable, I think. Which enzymes?, Plant tissue or animal? Cells
grown in culture?
I am sure someone on the list can help if your question is more
specific.

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (908)-235-4583; fax -4029 e-mail: mcauliff-at-.umdnj.edu
***************************************************************




From: I.Ivanov-at-ix.netcom.com
Date: Tue, 11 Mar 1997 11:54:41 -0600 (CST)
Subject: Re: Balzer's freeze-fracture

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On 03/10/97 17:21:17 you wrote:
} I would like tofind someone in the S.F. Bay area that has a freeze-fracture
} machine that would be available for doing samples. I have a colleague in
} the area who is looking for a facility where she might be able to do some
} samples.
} We would be very grateful for any leads. Thanks in advance. ML
}
} Mei Lie Wong
} Department of Biochemistry
} HHMI-UCSF
} Ph. 415-476-4441 Fax 415-476-1902
} email wong-at-msg.ucsf.edu

Dear Mei,
Please call Nancy Smith at Cal State Hayward (main number 510-885-3000). I
am sure she cal help you.
Regards

Igor C. Ivanov, SIMS,Auger,TEM,SEM
Senior Scientist Analytical Services
RJ LeeGroup, Inc. Contract Research
530 McCormick Str. (510)-567-0480 phone
San Leandro, CA 94577 (510)-567-0488 FAX





From: Donna Jacobs :      jacobs-at-uimrl7.mrl.uiuc.edu
Date: Tue, 11 Mar 1997 14:29:37 -0600
Subject: Re: Open Position - Research Electron Microscopist

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Return-Path: {das3-at-Lehigh.EDU}
} X-Sender: das3-at-mail.Lehigh.EDU
} Date: Tue, 11 Mar 1997 16:26:22 +0100
} To: Donna Jacobs {jacobs-at-uimrl7.mrl.uiuc.edu}
} From: das3-at-Lehigh.EDU (David A. Smith)
} Subject: Re: Open Position - Research Electron Microscopist
}
} DR. DAVID SMITH PASSED AWAY SEPTEMBER 1996. PLEASE REMOVE FROM ALL E-MAIL
} AND MAIL LISTINGS. THANK YOU.
}
} MAXINE MATTIE
}
} David A. Smith, Department of Materials Science and Engineering, Lehigh
} University, 5. E. Packer Ave., Bethelehem, PA 18015, Ph 610 758 4231, Fax
} 610 758 4244
}
}
}
Donna Jacobs
MRL Administration
University of Illinois
104 South Goodwin Avenue
Urbana, Illinois 61801
Phone (217) 244-2944
Fax (217) 244-2946
email - jacobs-at-uimrl7.mrl.uiuc.edu





From: Damian_Neuberger_at_RLT013-at-ccmailgw.mcgawpark.baxter.com
Date: Tue, 11 Mar 1997 14:49:52 -0600
Subject: Refractive Index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Good Afternoon (Central Std Time USA) all Microscopists:

I have just had a request to determine the refractive index of zinc
pyrithione, a crystalline, colorless, transparent powder. Does anyone
know what the RI is? It's not in McCrone Atlas. I will be getting a
sample Monday and with a set of Cargille RI liquids, I will determine
the RI.

Alternatively, I have a B & L Abbe type refractometer used to
determine refractive index of liquids. However, I thought I read
somewhere that it could be used to determine the RI of solids. If
this is in fact possible, does anyone out there know how?

TIA

Damian Neuberger
neuberd-at-baxter.com




From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Tue, 11 Mar 1997 14:48:36 -0800 (PST)
Subject: CPD Thanks!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Thanks to all who responded to my cpd inquiry. The consensus seems to be
that fresh, dry 100% ETOH is very important and that I need to keep my
samples longer in the liquid CO2 to get a better exchange with the ETOH.

I'll try out all the suggestions. Hopefully, the only raisins I'll get
from now on will be in my breakfast cereal. = )


Thanks again.


Paula = )


p.s. I'm going to phone some vendors of the resins & such to see if
they've done studies as to which type of gloves work best. I'll keep you
posted.

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu






From: James Ekstrom :      jekstrom-at-exeter.edu (by way of Nestor J. Zaluzec)
Date: Tue, 11 Mar 1997 18:01:04 -0500
Subject: ISI SEM Manual needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am acquiring an ISI SX-30E and need a copy of the manual for this
instrument without paying the hundreds of dollars Topcon wants for such
a manual.
Any suggestions where I might be able to get a xerox of the manual or
borrow a copy to xerox.
Any help would be appreciated.

Jim Ekstrom
Phillips Exeter Academy






From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Tue, 11 Mar 1997 15:08:48 -0400
Subject: RE:SIP Service

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Duniway Stockroom Corp. (800-446-8811; info-at-duniway.com) advertise a
service for rebuilding sputter ion pumps (p. 25 of their catalog), and
state that a properly rebuilt pump should "have the same performance as a
new pump". The cost of rebuilding varies from about $400 to about $3000,
depending on the size and type of pump
No commercial interest - just trying to be helpful.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Wed, 12 Mar 1997 17:04:19 +1200
Subject: Vascular casting - Mercox supplier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone who performs vascular casting know where to obtain a product
called Mercox CL2? A fellow electron microscopist has heard that it is the
latest thing in vascular casting when mixed with methyl methacrylate but
hasn't been able to get any further info.
If anyone knows who makes it and where it can be bought we would like to hear.

Thanks in advance

Richard Easingwood
South Campus Electron Microscope Unit
School of Medical Sciences
University of Otago
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

e-mail: richard.easingwood-at-stonebow.otago.ac.nz








From: Jim Darley :      jim-at-proscitech.com.au
Date: Wed, 12 Mar 1997 17:28:42 +1000
Subject: Re: Refractive Index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

For the theory you need to look up Snell's Law and Molecular refractivity.
The RI relates to the velocity of light through various substances. The
difference between air and water RI results in a stick appearing angled at
the interface.
Transparent specimen become invisible if immersed in a liquid of identical
RI. This is extensively used in police scientific work were glass fragments
can be identified to come from a crime scene. Very simple, put a sample
onto a microscope slide. Add a drop of RI solution. If the submersed
specimen becomes invisible under the microscope, than the refractive index
of the specimen is that of the RI test solutions. Some experimenting to
find the right solution is required. Good luck.
Cheers Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 77 740 370 Fax: +61 77 892 313
Great microscopy catalogue, 300+ Links, MSDS
************************ http://www.proscitech.com.au
----------------------------------------------------------------------------
-----
} From: Damian_Neuberger_at_RLT013-at-ccmailgw.mcgawpark.baxter.com
} To: microscopy-at-Sparc5.Microscopy.Com
} Subject: Refractive Index
} Date: Wednesday, 12 March 1997 6:49
}
} I have just had a request to determine the refractive index of zinc
} pyrithione, a crystalline, colorless, transparent powder. Does
anyone
} know what the RI is? It's not in McCrone Atlas. I will be getting
a
} sample Monday and with a set of Cargille RI liquids, I will
determine
} the RI.





From: Frank Karl :      fskarl-at-goodyear.com
Date: Wed, 12 Mar 1997 10:09:03 -0500
Subject: Re: Refractive Index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Transparent specimen become invisible if immersed in a liquid of identical
} RI. This is extensively used in police scientific work were glass fragments
} can be identified to come from a crime scene. Very simple, put a sample
} onto a microscope slide. Add a drop of RI solution. If the submersed
} specimen becomes invisible under the microscope, than the refractive index
} of the specimen is that of the RI test solutions.

Unfortunately it is not quite that simple. If your material is a glass
and/or has one refractive index
this procedure will work fine. Don't forget to correct for temperature and
wavelength!

But-- If your sample has two or more refractive indexes the procedure
becomes a lot more complicated. You need to separate one refractive index
from another and this can be done with the polarized light microscope with
a lot of practice. I suggest "Handbook of Chemical Microscopy, Vol One" by
Chamot and Mason, if you can find a copy, as a reference. The procedures
and techniques are too detailed for a short note.

Best wishes.... Frank

PS i was unable to find any optical data in Winchell's "The Optical
Properties of Organic Compounds"



----------------------------------------------------------------------------
-------

These opinions are mine alone and have no relationship to my employer.
Thank you.

Frank Karl fskarl-at- goodyear.com
Goodyear Tire & Rubber Co. Voice: 330-796-7818
Analytical Services Dept. 415B Fax: 330-796-3304
142 Goodyear Blvd.
Akron, OH 44305
USA

They that give up essential liberty to obtain a little temporary safety
deserve neither liberty or safety. Benjamin Franklin




From: Bob_Citron-at-cc.chiron.com (Bob Citron)
Date: 3/11/97 2:49 PM
Subject: Refractive Index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Damian;

It is possible to determine the RI of SOME solids with the B&L Abbe
refractometer. But I don't know if it is possible to do this on a powder;
I have never done it. The more critical requirements for doing a solid (if
you are still interested) are:

1) You must have very intimate contact between the solid and the lower prism.
This is normally accomplished by polishing the sample such that it is extremely
flat, and then a contact fluid (1-bromonapthalene) is used to form the intimate
contact. Use as little contact fluid as necessary (usually a very small drop).

2) It is important to have a 90 degree angle between the bottom prism and one
edge of your sample. This edge also needs to be highly polished, like the
contact surface.

Good luck with the powder; if anyone suggests a way to do this with the
refractometer, I would also be interested.

Regards,

-Bob
************************************
Bob Citron
Chiron Vision
555 W. Arrow Hwy
Claremont, CA 91711
ph: (909) 399-1311
email: Bob_Citron-at-cc.chiron.com
************************************

______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Good Afternoon (Central Std Time USA) all Microscopists:

I have just had a request to determine the refractive index of zinc
pyrithione, a crystalline, colorless, transparent powder. Does anyone
know what the RI is? It's not in McCrone Atlas. I will be getting a
sample Monday and with a set of Cargille RI liquids, I will determine
the RI.

Alternatively, I have a B & L Abbe type refractometer used to
determine refractive index of liquids. However, I thought I read
somewhere that it could be used to determine the RI of solids. If
this is in fact possible, does anyone out there know how?

TIA

Damian Neuberger
neuberd-at-baxter.com




From: beth-at-dogwood.botany.uga.edu (Beth Richardson)
Date: Wed, 12 Mar 1997 11:48:49 -0500
Subject: jobs?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199703121643.LAA26301-at-dogwood.botany.uga.edu}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hi,
I deleted the job posting from Auburn, AL (and I think there was one from
NC, too). If anyone saved them would they please send em to me.

Thanks very much,
Beth

**************************************
Beth Richardson
EM Lab Coordinator
Botany Department
University of Georgia
Athens, GA 30602

Phone - (706) 542-1790
FAX - (706) 542-1805
Email - beth-at-dogwood.botany.uga.edu
**************************************






From: Gary Dietrich Chinga :      garyc-at-stud.ntnu.no
Date: Wed, 12 Mar 1997 18:31:10 +0100 (MET)
Subject: Tonicity and osmolarity?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi!

Can someone explain the relation between the tonicity and osmolarity
of a fixation solution?


Gary.






From: Richard Thrift :      Richard_Thrift-at-depotech.com
Date: Wed, 12 Mar 1997 11:02:04 -0800
Subject: Tonicity and osmolarity -Reply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Osmolarity is an easily measurable characteristic, by e.g. freezing point
depression. In simple terms it is the total concentration of solutes
(including ions) in a solution, and this does not depend on whether the
solutes can cross cell membranes or not. Tonicity relates to the osmotic
GRADIENT due to solutes that affects a semi-permeable membrane ( i.e.
a membrane that is permeable to water); ONLY solutes that do not cross
the membrane contribute to this effect. The two characteristics are
different and obviously have vastly different consequences on
membranes. Weak bases, even though ionized, have some measurable
permeability and so their contribution to tonicity must be much less than
their contribution to osmolarity. Osmium is hydrophobic and quite
permeable through membranes, so contributes *nothing* to an osmotic
gradient that can disrupt the membranes (even assuming it doesn't alter
their permeability to tonic agents like sucrose). So osmium contributes
nothing to tonicity, but certainly does contribute to osmolarity.

I personally wonder how to express the tonicity of compounds that are
somewhat permeable, such as weak bases or short carboxilic acids &
aldehydes. I guess the concept of tonicity either doesn't apply or is
operational in such cases, depending on the time scale of interest. Can
anyone comment or give a reference to a good detailed textbook?
Something about reflection coefficients?

Richard

} } } Gary Dietrich Chinga {garyc-at-stud.ntnu.no} 03/12/97 09:31am } } }
Hi!

Can someone explain the relation between the tonicity and osmolarity of
a fixation solution?


Gary.








From: wise-at-vaxa.cis.uwosh.edu
Date: Wed, 12 Mar 1997 13:37:26 +0000
Subject: Re: Tonicity and osmolarity?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Gary,

Tonicity is a relative, unitless comparison of one solution to
another in which the first solution (presumably the water in your tissue
sample) is hypertonic, hypotonic or isotonic to the second solution
(presumably your fixative). Osmolarity is an absolute scale (usually in
some type of pressure unit) which describes the concentration of osmolytes
in a single solution. So you need to know the osmolarity of your tissue so
you can set the osmolarity of your fixative to be isotonic with respect to
the tissue.

Cheers

Bob


Robert R. Wise
Plant Physiologist and Director, UWO Electron Microscope Facility
Department of Biology
University of Wisconsin Oshkosh
Oshkosh, WI 54901
(414) 424-3404 tel
(414) 424-1101 fax
wise-at-uwosh.edu






From: Goodhouse, Joseph :      jgoodhouse-at-molecular.princeton.edu
Date: Wed, 12 Mar 97 14:56:00 PST
Subject: Cell-Tak

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear list,
Could someone please tell me where I can get Cell-Tak from or Muscle
Adhesion Protein.

Thank You.

J. Goodhouse




From: R. Edelmann :      edelmare-at-CASMAIL.MUOHIO.EDU
Date: Wed, 12 Mar 1997 15:29:20 -500
Subject: Denkil-SEM ?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have a user trying to follow a new protocol which uses "an
antistatic solution of ... Denkil-SEM (Hodogaya Chemical)" but I
have no idea what "Denkil-SEM" is can anyone out there give us a
hint? (Before we start trying other antistatic solutions, i.e. Static
Guard, or Cling Free!).

Thanks.


Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Supervisor
352 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513-529-5712 Fax: 513-529-4243
E-mail: edelmare-at-muohio.edu




From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Wed, 12 Mar 1997 11:26:19 -1000 (HST)
Subject: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have a question which has undoubtedly been answered before...! We are
going to buy a flat bed scanner, probably a Microtek with a transparency
adapter. We would primarily use it for scanning in TEM and SEM (polaroid)
negatives. Should we go for the 24-bit color, 300 x 600 dpi version, or
the more expensive (ouch) 30 bit color, 600 x 1200 dpi version? Our
resident computer guru says that we should go with the cheaper version
because we are unlikely to have an output source with better than 600 dpi
for most stuff. It's true; the printers around here are mostly 300 - 600
dpi. Is there a compelling reason for the better model? Will I be really
sorry in a year if I don't?

Thanks you all for all the expertise and advice on so many different
topics!

Aloha,
Tina

http://www.pbrc.hawaii.edu/bemf/microangela
****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************





From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Wed, 12 Mar 1997 14:29:53 -0800 (PST)
Subject: More glove results

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello Everybody,


I've been checking around on the glove issue. It seems that there
is no one glove that works for everything (darn!). Usually if it's good
for one thing, it's lousy for another.
We work with Lowicryl resins a lot and the best glove to use is........

Nitrile, those blue smelly things that don't stretch worth
a darn.

I also received some info. from a paper (thanks, Bill = ) ), that
told of an actual test done with various resins & such on different makes
of gloves.
The paper is "Glove Material For Handling Epoxy Resins" by David L. Ringo,
Douglas R. Read and Eugene H. Cote-Robles in the Journal of Electron
Microscopy Techniques 1:417-418 (1984). Pretty interesting stuff!

Safety in the lab is an important thing. So each lab must make
it's own mind as to which type of glove they want to use. The important
thing is that you get your students to glove up and work in the fume hoods
and be aware that almost all of the chemicals in an EM lab are hazardous.


Happy Scoping!


Paula = )

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu






From: pflaitz-at-vnet.ibm.com
Date: Wed, 12 Mar 97 17:13:18 EST
Subject: Meeting of Metropolitan Microscopy Society.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

*********************************************
* Metropolitan Microscopy Society Meeting *
*********************************************

Date: Wednesday, March 26, 1997

Time: 10:00 AM

Place: Howard Johnson Lodge, 393 Route 17, Paramus, NJ

Directions: The Howard Johnson Lodge is on the southbound lane of
Route 17. From the Garden State Parkway, use Exit 163
if you are northbound, and Exit 165 if you are
southbound.




****************************
* Agenda *
****************************


10:00 - 10:30 am Registration ($5.00), Coffee and Danish.


10:30 - 10:45 am Introductory remarks and society announcements --
Philip Flaitz.


10:45 - 11:30 am HIGH RESOLUTION LOW VOLTAGE SCANNING ELECTRON
MICROSCOPY, Dr. Frederic Cosandey, Dept. of Ceramic and Materials
Science, Rutgers, The State University of New Jersey, Piscataway,
NJ.

Scanning Electron Microscopy at low voltages (0.2 - 5 keV) has many
advantages over more conventional electron energies such as reduced
electron range, increased secondary electron yield, reduced
radiation damage and charging artifacts. However, in order to
maintain high resolution at the lowest voltages, special lens
designs are required to correct for the chromatic aberration of the
probe forming objective lens. In addition, improved secondary
electron detection systems must be implemented. In this talk, the
relative merits of various objective lens designs will be discussed
with special emphasis on the electrostatic lens with retarding
field concept. Also, the unique advantages of low voltage SEM for
secondary and backscattered electron imaging modes will be
discussed. A few specific examples will be presented including
domain size measurements in block co-polymers, oxide surface
structure determination and phase identification in nanocomposites.

11:30 - 12:15 pm MICROBEAM ANALYSIS OF POLYMERS, DRUGS AND INOR-
GANIC MOLECULES BY INFRARED MICROSPECTROSCOPY, Dr. John A. Reffner,
Spectra-Tech Inc., Shelton, CT 06484

Infrared microspectroscopy (IMS) is a growing technology for
microbeam analysis of organic and molecular materials. As the
fusion of scanning electron microscopy with x-ray emission
spectroscopy created an exciting way for microscopists to
investigate elemental composition, so has IMS impacted the way we
now investigate the molecular chemistry of materials. IMS is
essentially a photon probe --- producing spectral data by the
absorption or reflection of infrared radiation. Applications to
polymers, pharmaceuticals, electronics and forensics illustrate the
wide range of uses of IMS. In reviewing IMS technology, both the
strengths and limitations of current instruments are presented with
a look at directions for future developments.


12:30 - 1:30 pm Lunch.

For additional information, please contact:

Phil Flaitz, IBM Analytical Services
pflaitz-at-vnet.ibm.com, 914-892-3094





From: drstad-at-juno.com (David R Stadden)
Date: Wed, 12 Mar 1997 18:24:07 EST
Subject: Zerostat 3

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I know this question has been asked here before, so bear with me. Where
can I get one of those red antistatic guns, now that they are no longer
handled by Fullam? Their representative told me that there is supposedly
a hefty supply of these units still in the U.K., and that Fullam doesn't
handle them anymore because they'd have to buy more than they could sell.
I know Discwasher used to distribute them. Does anyone know of the
English company that has this huge supply? Are there any distributors?
Is there any similar product? Appreciate any leads.

Dave Stadden
DRStad-at-Juno.com




From: Donald Lovett :      lovett-at-tcnj.edu
Date: Wed, 12 Mar 1997 20:03:23 -0500 (EST)
Subject: Re: Tonicity and osmolarity?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Wed, 12 Mar 1997, Gary Dietrich Chinga wrote:

} Can someone explain the relation between the tonicity and osmolarity
} of a fixation solution?

= = = = = = = = = = = = = = = =

Osmolarity is a measure of one of the colligative properties (osmotic
pressure) of a solution. In a rough sense it is the measure of the amount
of solute in a solution, but the degree of dissociation of the solute
affects its osmolarity, so osmolarity is not necessarily a direct measure
of molar concentration. (For example, a 1 molar solution of NaCl will
have *approximately* twice the osmotic pressure [osmolarity] of a 1 molar
solution of sucrose, since the NaCl dissociates into 1 molar Na+ and
1 molar Cl-.)

Two solutions (not necessarily of the same solute or of a single solute)
are isosmotic if they have the same osmolarity (osmotic pressure). If two
isosmotic solutions are placed on opposite sides of a semipermeable
membrane, the osmotic pressure on each side is the same and there is no
*net* movement of water across the membrane. However, if the osmolarity
of the two solutions is not the same, then water will move across the
membrane from the hypo-osmotic solution (more dilute) to the hyperosmotic
solution (less dilute) until the osmotic pressure on each side is equal
(a number of other factors affect this as well, but let's ignore them
here).

Tonicity refers to the response of *cells or tissues* to the solutions in
which they are immersed. If cells are placed in a hypertonic solution,
net movement of water will be out of the cell, causing the cell to
shrivel. If cells are placed in a hypotonic solution, net movement of
water will be into the cell, causing the cell to swell or burst. Tonicity
is useful only in reference to a particular cell or tissue.

Thus, the microscopist wishes to add sufficient solute(s) to the fixation
solution so that the solution has the correct osmolarity (measured in
milliosmols) so that it will have the desired tonicity with respect to the
cells that are being exposed to the fixative. [There is debate whether
the solution should be slightly hypertonic or hypotonic, but then that is
another subject about which we may choose to debate.]

To summarize, osmolarity is a measurement of solute concentration
(measurement can be made in a beaker). Tonicity is a comparison of
osmolarities between a cell and the solution to which it is exposed.

I hope that this does not leave readers more confused than they were
before.
}

______________________________________________________________________
Donald L. Lovett e-mail: lovett-at-tcnj.edu
Assoc. Professor, Dept. of Biology voice: (609) 771-2876
The College of New Jersey fax: (609) 771-2674
Trenton, NJ 08650-4700







From: WARRENJ1-at-cliffy.polaroid.com
Date: 3.12.97 4:26 PM
Subject: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The 30 bits will give you additional shadow and highlight detail that
24 bits will not provide. Now, you may ask 'my software can only
handle 24 bits so what good does 30 bits do when I will have to
convert it down?'. When you convert the file to 24 bits, the system
will choose from the best 24 bits to produce the converted file. I
would equate it to a survey or poll - the larger your sample, the more
accurate your results.

John D. Warren
Area Sales Manager
Digital Products
Polaroid Corporation



______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I have a question which has undoubtedly been answered before...! We are
going to buy a flat bed scanner, probably a Microtek with a transparency
adapter. We would primarily use it for scanning in TEM and SEM (polaroid)
negatives. Should we go for the 24-bit color, 300 x 600 dpi version, or
the more expensive (ouch) 30 bit color, 600 x 1200 dpi version? Our
resident computer guru says that we should go with the cheaper version
because we are unlikely to have an output source with better than 600 dpi
for most stuff. It's true; the printers around here are mostly 300 - 600
dpi. Is there a compelling reason for the better model? Will I be really
sorry in a year if I don't?

Thanks you all for all the expertise and advice on so many different
topics!

Aloha,
Tina

http://www.pbrc.hawaii.edu/bemf/microangela
****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************




From: Warren Straszheim :      wes-at-ameslab.gov
Date: Wed, 12 Mar 1997 17:32:16 -0600
Subject: Re: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would suggest to check out the speed as well as the resolution. We have an
old HP-IIcx here that easily beats a newer cheaper scanner hands down. The
software is also not as convenient as HP's deskscan package.

Regarding resolution, remember it will take a few of those printer pixels to
dither up the gray scale for each of your digitized pixels. Thus, a 600 dpi
printer may only be able to handle about a 200 dpi image. Of course if you
are scanning 35 mm slides and enlarging them, you will need the high res
scanning.

At 11:26 AM 3/12/97 -1000, Tina wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

E-Mail: wes-at-ameslab.gov (or: wesaia-at-iastate.edu)
http://www.public.iastate.edu/~iprt_info/cfce/ (re: coal)
http://www.public.iastate.edu/~wesaia/marl/ (re: SEM)

coal characterization and processing
electron microscopy, x-ray analysis, image analysis
computer applications





From: Richard Thrift :      Richard_Thrift-at-depotech.com
Date: Wed, 12 Mar 1997 22:09:42 -0800
Subject: Tonicity and osmolarity -Reply II (a vote for empiricism,

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Donald Lovett's explanation quoted below is very good except that it
perpetuates the myth that tonicity is related to the difference between
two OSMOLARITIES on opposite sides of the membrane. That is ONLY
true if the membrane involved is impermeable to ALL of the solutes
contributing to the osmolarity. Since formaldehyde is soluble in both
benzene and chloroform I expect it to be quite permeable across cell
membranes. Thus it will contribute to osmolarity, but should contribute
little to tonicity. The tonicity of your fixative should be adjusted with other
components, and the final osmolarity of an isotonic fixative will be approx
300 mOsm PLUS the value of osmolarity contributed by the formaldehyde
present.

To reiterate, IF a compound is permeable across the membrane it
contributes to osmolarity but does not contribute to the osmotic
GRADIENT. Thus it does NOT contribute to tonicity.

Caveats: Practically speaking, the rate at which a permeable solute
crosses the membrane will affect the transient forces on the membrane.
I assume tonicity is ill-defined unless the system is at equilibrium. The
situation becomes even more complicated if the compounds added (such
as aldehydes or alcohols) alter the permeability of the membrane for
other buffer components present. And if differences in rates of
diffusion through a chunk of tissue are involved, are we really still talking
about tonicity?? That calls for empirical, not theoretical, optimization of
the recipe used !!

On that note, how many of you microscopists have actually compared
e.g. confocal images of cell cultures fixed with buffers of different
tonicity with equivalent images of live cultured cells? The time to diffuse
through a monolayer of cells should be negligible. What do you find is
optimum?

Richard


} } } Donald Lovett {lovett-at-tcnj.edu} 03/12/97 05:03pm } } }
Osmolarity is a measure of one of the colligative properties (osmotic
pressure) of a solution. In a rough sense it is the measure of the amount
of solute in a solution, but the degree of dissociation of the solute affects
its osmolarity, so osmolarity is not necessarily a direct measure of molar
concentration. (For example, a 1 molar solution of NaCl will have
*approximately* twice the osmotic pressure [osmolarity] of a 1 molar
solution of sucrose, since the NaCl dissociates into 1 molar Na+ and 1
molar Cl-.)

Two solutions (not necessarily of the same solute or of a single solute)
are isosmotic if they have the same osmolarity (osmotic pressure). If
two isosmotic solutions are placed on opposite sides of a semipermeable
membrane, the osmotic pressure on each side is the same and there is
no
*net* movement of water across the membrane. However, if the
osmolarity of the two solutions is not the same, then water will move
across the membrane from the hypo-osmotic solution (more dilute) to the
hyperosmotic solution (less dilute) until the osmotic pressure on each
side is equal
(a number of other factors affect this as well, but let's ignore them here).

Tonicity refers to the response of *cells or tissues* to the solutions in
which they are immersed. If cells are placed in a hypertonic solution, net
movement of water will be out of the cell, causing the cell to shrivel. If
cells are placed in a hypotonic solution, net movement of water will be
into the cell, causing the cell to swell or burst. Tonicity is useful only in
reference to a particular cell or tissue.

Thus, the microscopist wishes to add sufficient solute(s) to the fixation
solution so that the solution has the correct osmolarity (measured in
milliosmols) so that it will have the desired tonicity with respect to the
cells that are being exposed to the fixative. [There is debate whether the
solution should be slightly hypertonic or hypotonic, but then that is
another subject about which we may choose to debate.]

To summarize, osmolarity is a measurement of solute concentration
(measurement can be made in a beaker). Tonicity is a comparison of
osmolarities between a cell and the solution to which it is exposed.

I hope that this does not leave readers more confused than they were
before.
}

______________________________________________________________________
Donald L. Lovett e-mail: lovett-at-tcnj.edu
Assoc. Professor, Dept. of Biology voice: (609) 771-2876
The College of New Jersey fax: (609) 771-2674
Trenton, NJ 08650-4700









From: wahlbrin-at-crpcu.lu (Petra Wahlbring)
Date: Thu, 13 Mar 97 09:26 MET
Subject: Post-doc position: analytical TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I'd like to forward this announcment of an open post-doc position to the
list:


} The Laboratoire d'Analyse des Materiaux (LAM), of the "Centre de Recherche
} Public-Centre Universitaire" in Luxembourg, has an immediate opening for a
} two years post-doc position.
} The core of the subject to be covered deals with analytical TEM (EELS and
} EDX) : sensitivity, quantification, artifacts,... on complex samples.
} As a side subject, this person will have to help in optimizing the
} preparation of cross sectional samples: ion dimpling/microtome sectioning.
}
} If you have the necessary qualifications, please send your resume to :
} Dr. H.N. Migeon
} Director of LAM
} 162a, Avenue de la Faiencerie
} L-1511 Luxembourg
} After a first selection, interviews will take place in Luxembourg. No
} funding will be provided for overseas trips.
}
} Henri-Noel Migeon


Best regards,

Petra

--------------------------------------------------------------
Dr. Petra Wahlbring
Centre de Recherche Public Centre Universitaire (CRP-CU)
Laboratoire d'Analyse des Materiaux (LAM)
162a, av. de la Faiencerie L-1511 Luxembourg
tel. +352-466644-402 fax +352-466644-400
e-mail: petra.wahlbring-at-crpcu.lu or 100112.2335-at-compuserve.com





From: Andrey L. Chuvilin :      dusha-at-catalysis.nsk.su
Date: Thu, 13 Mar 1997 14:39:25 +0700 (GMT)
Subject: Re: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Wed, 12 Mar 1997 WARRENJ1-at-cliffy.polaroid.com wrote:

} The 30 bits will give you additional shadow and highlight detail that
} 24 bits will not provide. Now, you may ask 'my software can only
} handle 24 bits so what good does 30 bits do when I will have to
} convert it down?'. When you convert the file to 24 bits, the system
} will choose from the best 24 bits to produce the converted file. I
} would equate it to a survey or poll - the larger your sample, the more
} accurate your results.

This statement seems a little bit tricky.
24 color bits gives one 256 grey shadows.
If one want only to store and reproduce images - this is more than enough
(128 is O'K for viewing).
If one want to quantify grey levels for processing, 256 levels is also
usually enough because of narrow linear range of film, grain, etc.
Interpolation will give better results.
If one want to waste money - I have no reasons not to this :)





From: wa5ekh :      wa5ekh-at-cyberramp.net
Date: Thu, 13 Mar 1997 01:45:24 -0600
Subject: Wanted: Vacuum Coating System and Balzers 360

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Need to find a floor model lab vacuum coating systemwith : Defussion pump,
HV(2000 Volt) 200 ma. feed thru, low voltage high current feed thru,
cryotrap,etc. Also looking for an older Balzers 360 Freeze Fracture
system(any condition) and an old hummer or a sputtering head for Au-Pd.
By the way does anyone keep up with the "Particle-Protein" Freeze
Fracture theories and can you send me some references? Been a while.

Jeff Day
3208 Statler
Mesquite, Texas 75150
Day Phone:972-975-4338





From: Richard Beanland +44 1327 356363 :      richard.beanland-at-gecm.com
Date: Thu, 13 Mar 1997 11:23:52 +0000 (GMT)
Subject: TEM: Fresnel fringes from amorphous layers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello all,
I have been looking at very thin (~10-30 A) multilayers of silicon
oxide and nitride. I have very simplistically assumed that the position of
the interface lies half-way between the first bright and dark fringe on each
side of the interface. Does anyone know whether this is a valid assumption -
and if there's any software that I can use to simulate the image? (Or, failing
this, where I can get hold of the theory to let me do this myself).

Many thanks in advance,

Richard Beanland,
GMMT Ltd.,
Caswell,
Towcester,
Northants NN12 8EQ
UK





From: Bill Miller :      microbill-at-MOHAWK.NET
Date: Thu, 13 Mar 1997 07:31:19 -0500
Subject: Re: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Tina Carvalho wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} I have a question which has undoubtedly been answered before...! We are
} going to buy a flat bed scanner, probably a Microtek with a transparency
} adapter. We would primarily use it for scanning in TEM and SEM (polaroid)
} negatives. Should we go for the 24-bit color, 300 x 600 dpi version, or
} the more expensive (ouch) 30 bit color, 600 x 1200 dpi version? Our
} resident computer guru says that we should go with the cheaper version
} because we are unlikely to have an output source with better than 600 dpi
} for most stuff. It's true; the printers around here are mostly 300 - 600
} dpi. Is there a compelling reason for the better model? Will I be really
} sorry in a year if I don't?
}
} Thanks you all for all the expertise and advice on so many different
} topics!
}
} Aloha,
} Tina
}
} http://www.pbrc.hawaii.edu/bemf/microangela
} ****************************************************************************
} * Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
} * Biological Electron Microscope Facility * (808) 956-6251 *
} * University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
} ****************************************************************************
Tina - buy the most true optical resolution and pixel depth you can
afford - not so much for the SEM stuff but for the TEM negatives. The
scanner should have enough optical (not interpolated) resolution to over
sample the film resolution by at least a factor of 2.5 to 3 (the Nyquist
sampling limit you know) otherwise you will not be able to treat the
digitized images like the original negative - i.e. information will be
lost! The company that makes tha scanner probably also makes a 1k x 2k
version which would be even better.

Best Regards - Bill Miller




From: Donald Lovett :      lovett-at-tcnj.edu
Date: Thu, 13 Mar 1997 08:37:47 -0500 (EST)
Subject: Re: Tonicity and osmolarity -Reply II (a vote for empiricism,

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Wed, 12 Mar 1997, Richard Thrift wrote:

}
} Donald Lovett's explanation quoted below is very good except that it
} perpetuates the myth that tonicity is related to the difference between
} two OSMOLARITIES on opposite sides of the membrane. That is ONLY
} true if the membrane involved is impermeable to ALL of the solutes
} contributing to the osmolarity. ......



} Richard:

Thanks for your extended clarification of the subject. I agree entirely
with your comments. I had tried to focus on the difference between
the two terms and simplify my response with the caveat:
}
} "(a number of other factors affect this as well, but let's ignore them
here)."
}
I also agree that the response of the cell to the solution (as evaluated
by trial and error) is the most important aspect to microscopists,
irrespective of how one names or measures the composition of the solution.


Don
}
} ______________________________________________________________________
} Donald L. Lovett e-mail: lovett-at-tcnj.edu
} Assoc. Professor, Dept. of Biology voice: (609) 771-2876
} The College of New Jersey fax: (609) 771-2674
} Trenton, NJ 08650-4700
}
}
}
}
}
}

______________________________________________________________________
Donald L. Lovett e-mail: lovett-at-tcnj.edu
Assoc. Professor, Dept. of Biology voice: (609) 771-2876
The College of New Jersey fax: (609) 771-2674
Trenton, NJ 08650-4700







From: WARRENJ1-at-cliffy.polaroid.com
Date: 3.13.97 2:39 AM
Subject: Re: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have found that what is 'adequate' for one opportunity may not be
adequate for another. In any event, for those who are unsure of a
decision between digital products based on varying specifications, I
suggest you look at the end result. Take an image and scan it in on a
24 bit scanner and a 30 bit scanner, print them out on the printer(s)
you would typically use and compare the results. If the 24 bit image
is sufficient for your application, then buy the thing. The street
price should be about $220. If you like the 30 bit converted to 24
better, then buy it. Its street price should be around $525.

As far as the comment regarding scanning 35 mm slides, I would not
recommend using a flatbed scanner to scan 35mm images that will be
enlarged more than 1x the original size, otherwise the image gets soft
even with a 600 dpi scanner. If you are scanning 35mm on a regular
basis, you should use a film scanner.

John D. Warren
Area Sales Manager
Digital Products
Polaroid Corporation


______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

On Wed, 12 Mar 1997 WARRENJ1-at-cliffy.polaroid.com wrote:

} The 30 bits will give you additional shadow and highlight detail that
} 24 bits will not provide. Now, you may ask 'my software can only
} handle 24 bits so what good does 30 bits do when I will have to
} convert it down?'. When you convert the file to 24 bits, the system
} will choose from the best 24 bits to produce the converted file. I
} would equate it to a survey or poll - the larger your sample, the more
} accurate your results.

This statement seems a little bit tricky.
24 color bits gives one 256 grey shadows.
If one want only to store and reproduce images - this is more than enough
(128 is O'K for viewing).
If one want to quantify grey levels for processing, 256 levels is also
usually enough because of narrow linear range of film, grain, etc.
Interpolation will give better results.
If one want to waste money - I have no reasons not to this :)




From: beth-at-dogwood.botany.uga.edu (Beth Richardson)
Date: Thu, 13 Mar 1997 09:33:35 -0500
Subject: thanks - job info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199703131428.JAA05812-at-dogwood.botany.uga.edu}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Hey, y'all
Just wanted to thank everyone who replied to my request for recent job
postings. Lots of people went digging through their trash and came up with
the information that I needed.
Thanks very much,
Beth

**************************************
Beth Richardson
EM Lab Coordinator
Botany Department
University of Georgia
Athens, GA 30602

Phone - (706) 542-1790
FAX - (706) 542-1805
Email - beth-at-dogwood.botany.uga.edu
**************************************






From: jm Lett :      jm_lett-at-cidmac.wustl.edu
Date: 13 Mar 1997 08:35:41 -0500
Subject: Re: Zerostat 3

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

RE: David R. Stadden's inquiry (Where can I get one of those red antistatic
guns, now that they are no longer handled by Fullam? Does anyone know of the
English company that has this huge supply? Are there any distributors? Is
there any similar product?)

I hope this is useful: Sigma Chemical Company (St. Louis, Missouri) carries
the Zerostat 3 (approx $55; US catalog #Z10,881-2). Their UK contact info
(from the list in the US catalog) is as follows:

Sigma-Aldrich Company Ltd.
Fancy Road, Poole,
Dorset, BH12 4QH
Free Tel: 0800 373731
Free Fax: 0800 378785
Tel: 01202 733114
Fax: 01202 715460

Jaclynn M. Lett
jm_Lett-at-cidmac.wustl.edu








From: Robert Underwood :      underwoo-at-u.washington.edu
Date: Thu, 13 Mar 1997 07:11:26 -0800 (PST)
Subject: Tonicity and Sucrose?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi

In reguards to the question of tonicity, does anyone know the best
percentage of sucrose to have in a primary Zamboni fixative(2% para and
15% picric acid in sorensons) to fix cell cultures in order to minimize
any cell distortion.

Thanks in advance

Bob Underwood
Morphology core
University of Washington





From: Robert Underwood :      underwoo-at-u.washington.edu
Date: Thu, 13 Mar 1997 07:11:26 -0800 (PST)
Subject: Tonicity and Sucrose?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi

In reguards to the question of tonicity, does anyone know the best
percentage of sucrose to have in a primary Zamboni fixative(2% para and
15% picric acid in sorensons) to fix cell cultures in order to minimize
any cell distortion.

Thanks in advance

Bob Underwood
Morphology core
University of Washington





From: Kalman Rubinson :      rubinsnk-at-is2.nyu.edu
Date: Thu, 13 Mar 1997 10:16:29 -0500 (EST)
Subject: Re: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Wed, 12 Mar 1997, Tina Carvalho wrote:

} Our resident computer guru says that we should go with the cheaper
} version because we are unlikely to have an output source with better
} than 600 dpi for most stuff. It's true; the printers around here are
} mostly 300 - 600 dpi. Is there a compelling reason for the better
} model? Will I be really sorry in a year if I don't?

One reason to go with the more expensive one is that the 30bit depth will
get you more contrast discrimination. Another reason is that, even with
300-600dpi printers, you might want to select a subregion of the scan and
expand it to full page. In that case, the extra resolution of the better
scanner will show.

Kal






From: ebs-at-ebsciences.com
Date: Thu, 13 Mar 1997 11:00:06 EST
Subject: Mercox resin

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Microscopists,

At 05:04 PM 3/12/97 +1200, Richard Easingwood wrote:
} Does anyone who performs vascular casting know where to obtain a product
} called Mercox CL2? A fellow electron microscopist has heard that it is the
} latest thing in vascular casting when mixed with methyl methacrylate but
} hasn't been able to get any further info.
} If anyone knows who makes it and where it can be bought we would like to hear.

There is a "Mercox Resin" in the Ladd catalog, but it doesn't say whether
it's CL2.

Best regards,
Steven Slap
********************************
Energy Beam Sciences, Inc.
Adding Brilliance To Your Vision
ebs-at-ebsciences.com
http://www.ebsciences.com/
********************************





From: Kim Rensing :      krensing-at-uvic.ca
Date: Thu, 13 Mar 1997 10:33:45 -0800
Subject: wet formvar?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Can formvar powder absorb water and cause holes in grid coatings, or is it
just the solvent in which it is made that causes the problems? We have some
rather old powder here and I wonder if it is still useable.

Thanks.

Kim Rensing






From: Peggy Bisher :      peggy-at-research.nj.nec.com
Date: Thu, 13 Mar 1997 15:05:08 -0400
Subject: Ceramic Grids/Rings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am looking for 3.05mm rings that I can use as a TEM grid made of
ceramic material. Does anybody know where I can find such thing?


Thank you, Peggy Bisher



NEC Research Institute
4 Independence Way
Princeton, NJ 08540.
Tel.: (609) 951-2629
Fax: (609) 951-2496
e-mail: peggy-at-research.nj.nec.com


Being defeated is often a temporary condition. Giving up is what makes it
permanent. -Marilyn vos Savant







From: cook-at-aaem.amc.anl.gov (Russell E. Cook)
Date: Thu, 13 Mar 1997 15:41:27 -0600
Subject: Re: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Tina,

We recently purchased a film scanner to digitize images and diffraction
patterns from 3.25"x4" negatives. Prior to this we had an old Microtek
flatbed with a transparency adapter (300 dpi). It was ok as long as we
didn't try to enlarge the image much.

I don't know how you will be using the digitized images. If all you want
to do is digitize the negatives and then print the images without enlarging
them, you don't need a high-resolution scanner. But I would recommend that
you get a scanner with a larger bit-depth.

We often need to do some significant image analysis, so we needed a scanner
which would give us very high optical resolutions and more than the usual
256 gray scale (24-bit color). We wanted at least 12 bits per color and
2400 dpi optical resolution. Our requirements may be more than you need.

I think John Warren's recommendation is good: try before you buy.

Russell E. Cook
Scientific Associate
Electron Microscopy Center for Materials Research
Argonne National Laboratory
Building 212
9700 South Cass Avenue
Argonne, IL 60439
(630)252-7194
FAX: (630)252-4798






From: mgb-at-nucleus.ansto.gov.au (Mark Blackford)
Date: Fri, 14 Mar 1997 08:44:01 +1200
Subject: SIP service

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

thanks to all who answered (either directly or via the listserver) my
request for info on SIP refurbishment. I'll summarise these responses and
make it available for anyone who contacts me directly. If there is enough
interest I'll post the summary on the listserver. Cheers,

Mark Blackford
TEM Group
Materials Division, ANSTO
PMB 1,
Menai, N.S.W.
Australia
2234
Phone 61 2 9717 3027
Fax 61 2 9543 7179

Disclaimer:
The views expressed in this E-mail message do not necessarily represent the
official views of ANSTO from which this message was conveyed.






From: Diana van Driel :      dianavd-at-eye.usyd.edu.au
Date: Fri, 14 Mar 1997 10:54:33 +1100
Subject: What resolution scanner?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Always get the best you can afford. And, yes, you will be sorry if you
don't! One of the uses I have found for the high resolution end of the
scanner is to check small details (eg is that cell junction really tight?)
in an EM negative without the hassle of a large print - it's amazing just
how much it is possible to see on the computer in a fraction of the time. I
have the ScanMaker 3 and am very happy with it. BUT, it is no good for
35mm, only prints and large format negatives.


Diana van Driel
Dept Ophthalmology
Sydney University C09
AUSTRALIA 2006






From: Diana van Driel :      dianavd-at-eye.usyd.edu.au
Date: Fri, 14 Mar 1997 11:32:33 +1100
Subject: Tonicity and Sucrose?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} In reguards to the question of tonicity, does anyone know the best
} percentage of sucrose to have in a primary Zamboni fixative(2% para and
} 15% picric acid in sorensons) to fix cell cultures in order to minimize
} any cell distortion.


I can't give a specific answer, but check out Maser MD et al: Relationships
among pH, osmolality and concentration of fixative solutions, Stain
Technology 42:175-182 (1967). You need to know what osmolality you want in
the fixative solution. Around 300 milliosmols (range 200-400) seemed to be
most common when I had to work it out too many years ago. In general the
contribution of the fixative can be ignored. So, work out the osmolality of
the buffer and add sucrose to bring the osmolality up to the total. For
instance, Sorensen's is approx 105 at 0.05M, approx 210 at 0.1M. Sucrose is
obvious (eg 0.2M = 200). It may be an idea to add some CaCl2 or MgCl2.

Diana van Driel
Dept Ophthalmology
Sydney University C09
AUSTRALIA 2006






From: Melvyn Dickson :      M.Dickson-at-unsw.edu.au
Date: Fri, 14 Mar 1997 11:39:34 +1000
Subject: Re: wet formvar?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Best practice is to dry the powder at 60 deg.C for half an hour just before
mixing it into the solvent. As for the solvent you ought to fractionate it
first. Redistil it and monitor the temperature of the vapour at the
stillhead and discard the (watery) fraction that comes off before the
correct boiling point of the solvent is reached. Stop the still when there
is around an (oily) tenth remaining of the original solvent.
}






From: Jim Darley :      jim-at-proscitech.com.au
Date: Fri, 14 Mar 1997 11:43:00 +1000
Subject: Re: Zerostat 3

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We supply the Zerostat, but in the US you could get one from Sigma.
Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 77 740 370 Fax: +61 77 892 313
Great microscopy catalogue, 300+ Links, MSDS
************************ http://www.proscitech.com.au
}


I know this question has been asked here before, so bear with me. Where
} can I get one of those red antistatic guns, now that they are no longer
} handled by Fullam? Their representative told me that there is supposedly
} a hefty supply of these units still in the U.K., and that Fullam doesn't
} handle them anymore because they'd have to buy more than they could sell.
} I know Discwasher used to distribute them. Does anyone know of the
} English company that has this huge supply? Are there any distributors?
} Is there any similar product? Appreciate any leads.
}
} Dave Stadden
} DRStad-at-Juno.com




From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Thu, 13 Mar 97 20:29:45 -0500
Subject: Re: Ceramic Grids/Rings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Peggy Bisher wrote:
=================================================
I am looking for 3.05mm rings that I can use as a TEM grid made of
ceramic material. Does anybody know where I can find such thing?
==================================================
About the closest thing about which I have any awareness would be our
diamond grids and rings, 3 mm diameter. They are sized to fit into any TEM
that takes a 3 mm grid. More information can be found about them on our
website.

Disclaimer: These are products offered by SPI Supplies and we would stand to
benefit if more people were using these diamond products.

Chuck


===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: http://www.2spi.com
############################
==================================================




From: mgb-at-nucleus.ansto.gov.au (Mark Blackford)
Date: Fri, 14 Mar 1997 17:25:34 +1200
Subject: SIP summary

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

As several people have expressed an interest in a summary of responses to
my SIP query I decided to post it to the listserver. My original posting
was:

} I am about to replace the sputter ion pump elements in our two JEOL TEM's.
} JEOL used to be able to recondition the elements used in their 2000FX
} microscopes but no longer provide this service. Can anyone suggest another
} company that does this sort of work? How long can I expect a reconditioned
} unit to last compared with a new one and what sort of cost would be
} reasonable given that JEOL are charging $2.5k and $4.5k (Australian
} dollars) for new units? Thanks in advance for any advice you can give.

Several people recommended contacting the following company:

DUNIWAY STOCKROOM Corp.
800-446-8811 also 415-969-8811 fax 415-965-0764
1600 N. Shoreline Blvd. Mountain View CA 94043
also 1305 Space Park Way,Mountain View,CA 94043
info-at-duniway.com

Other companies mentioned were:

Brechtel Mfg
Fremont,CA
510-732-9723 fax 510-732-9153

Vacuum Scientific Services
44 Ellesmere St, manchester M15 4JY UK
Tel. 44 161 833 9108
fax. 44 161 835 1443


Some of the comments received were:

A rebuilt pump is as good as a new one and should last as long.
They open up the pump, clean it out, replace all the elements,
and re-weld the case together. Prices are significantly less than
buying new.


Duniway has a rebuilding service and also sell their own brand of
pumps.
They state that rebuilt pumps have the same performance as a new pump.
If the pump has to be cut open to rebuild it, it can normally can be
rebuilt 2 or 3 times. They also say that their pumps can normally be
rebuilt 5-10 times. Being cut open is normally the case for small pumps.
On large pumps the titanium elements can often be removed through the
inlet port, allowing them to be rebuilt an unlimited number of times.

I emailed Duniway Stockroom Corporation and the following is an extract
from the reply by Eroc Inman:

We do rebuild not only the elements for JEOL pumps, but JEOL actually
sends all of their ion pumps to us for a complete overhaul/rebuild.

It your elements are just a standard size and configuration, like that
of Varian ion pumps, then the pricing is quite easy to determine. The
150 l/s pump elements are generally US$390.00/each for rebuilding. This
is the price for diode or triode configuration. I would be able to give
you an absolute price once we had the opportunity to take a look at
them. For now, I am quite certain that your elements are
not any different than what we generally see.


I have not yet contacted the other companies. If anyone can supply an
email address for them I would greatly appreciate it. I don't pretend that
this is a comprehensive list of companies providing SIP refurbishment, nor
do I recommend or have any affiliations with any one of them. Cheers,

Mark Blackford
TEM Group
Materials Division, ANSTO
PMB 1,
Menai, N.S.W.
Australia
2234
Phone 61 2 9717 3027
Fax 61 2 9543 7179

Disclaimer:
The views expressed in this E-mail message do not necessarily represent the
official views of ANSTO from which this message was conveyed.






From: Jim Darley :      jim-at-proscitech.com.au
Date: Wed, 12 Mar 1997 17:28:42 +1000
Subject: Re: Refractive Index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

For the theory you need to look up Snell's Law and Molecular refractivity.
The RI relates to the velocity of light through various substances. The
difference between air and water RI results in a stick appearing angled at
the interface.
Transparent specimen become invisible if immersed in a liquid of identical
RI. This is extensively used in police scientific work were glass fragments
can be identified to come from a crime scene. Very simple, put a sample
onto a microscope slide. Add a drop of RI solution. If the submersed
specimen becomes invisible under the microscope, than the refractive index
of the specimen is that of the RI test solutions. Some experimenting to
find the right solution is required. Good luck.
Cheers Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 77 740 370 Fax: +61 77 892 313
Great microscopy catalogue, 300+ Links, MSDS
************************ http://www.proscitech.com.au
----------------------------------------------------------------------------
-----
} From: Damian_Neuberger_at_RLT013-at-ccmailgw.mcgawpark.baxter.com
} To: microscopy-at-Sparc5.Microscopy.Com
} Subject: Refractive Index
} Date: Wednesday, 12 March 1997 6:49
}
} I have just had a request to determine the refractive index of zinc
} pyrithione, a crystalline, colorless, transparent powder. Does
anyone
} know what the RI is? It's not in McCrone Atlas. I will be getting
a
} sample Monday and with a set of Cargille RI liquids, I will
determine
} the RI.

#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: I.Ivanov-at-ix.netcom.com
Date: Tue, 11 Mar 1997 11:54:41 -0600 (CST)
Subject: Re: Balzer's freeze-fracture

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

On 03/10/97 17:21:17 you wrote:
} I would like tofind someone in the S.F. Bay area that has a freeze-fracture
} machine that would be available for doing samples. I have a colleague in
} the area who is looking for a facility where she might be able to do some
} samples.
} We would be very grateful for any leads. Thanks in advance. ML
}
} Mei Lie Wong
} Department of Biochemistry
} HHMI-UCSF
} Ph. 415-476-4441 Fax 415-476-1902
} email wong-at-msg.ucsf.edu

Dear Mei,
Please call Nancy Smith at Cal State Hayward (main number 510-885-3000). I
am sure she cal help you.
Regards

Igor C. Ivanov, SIMS,Auger,TEM,SEM
Senior Scientist Analytical Services
RJ LeeGroup, Inc. Contract Research
530 McCormick Str. (510)-567-0480 phone
San Leandro, CA 94577 (510)-567-0488 FAX

#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Tue, 11 Mar 1997 14:48:36 -0800 (PST)
Subject: CPD Thanks!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thanks to all who responded to my cpd inquiry. The consensus seems to be
that fresh, dry 100% ETOH is very important and that I need to keep my
samples longer in the liquid CO2 to get a better exchange with the ETOH.

I'll try out all the suggestions. Hopefully, the only raisins I'll get
from now on will be in my breakfast cereal. = )


Thanks again.


Paula = )


p.s. I'm going to phone some vendors of the resins & such to see if
they've done studies as to which type of gloves work best. I'll keep you
posted.

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu


#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Tue, 11 Mar 1997 15:08:48 -0400
Subject: RE:SIP Service

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Duniway Stockroom Corp. (800-446-8811; info-at-duniway.com) advertise a
service for rebuilding sputter ion pumps (p. 25 of their catalog), and
state that a properly rebuilt pump should "have the same performance as a
new pump". The cost of rebuilding varies from about $400 to about $3000,
depending on the size and type of pump
No commercial interest - just trying to be helpful.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321


#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Tue, 11 Mar 1997 11:40:21 -0800
Subject: re:staining enzymes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Dear Dr. Nunes:

Your question, "how to stain enzymes for EM" is so broad as to be
unanswerable, I think. Which enzymes?, Plant tissue or animal? Cells
grown in culture?
I am sure someone on the list can help if your question is more
specific.

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (908)-235-4583; fax -4029 e-mail: mcauliff-at-.umdnj.edu
***************************************************************
#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Wed, 12 Mar 1997 17:04:19 +1200
Subject: Vascular casting - Mercox supplier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Does anyone who performs vascular casting know where to obtain a product
called Mercox CL2? A fellow electron microscopist has heard that it is the
latest thing in vascular casting when mixed with methyl methacrylate but
hasn't been able to get any further info.
If anyone knows who makes it and where it can be bought we would like to hear.

Thanks in advance

Richard Easingwood
South Campus Electron Microscope Unit
School of Medical Sciences
University of Otago
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

e-mail: richard.easingwood-at-stonebow.otago.ac.nz




#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Tue, 11 Mar 1997 15:08:48 -0400
Subject: RE:SIP Service

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Duniway Stockroom Corp. (800-446-8811; info-at-duniway.com) advertise a
service for rebuilding sputter ion pumps (p. 25 of their catalog), and
state that a properly rebuilt pump should "have the same performance as a
new pump". The cost of rebuilding varies from about $400 to about $3000,
depending on the size and type of pump
No commercial interest - just trying to be helpful.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321


#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Fri, 14 Mar 1997 09:48:12 -0800
Subject: Re:tonicity of fixatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

In the theme of Diana van Driel's post check out Arborgh et al. "The
osmotic effect of glutaraldehyde during fixation". J. Ultrastr. Res.
56:339-350, 1976. A very interesting and through study.

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (908)-235-4583; fax -4029 e-mail: mcauliff-at-umdnj.edu
***************************************************************




From: Damian_Neuberger_at_RLT013-at-ccmailgw.mcgawpark.baxter.com
Date: Fri, 14 Mar 1997 09:04:37 -0600
Subject: refractive index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Thanks to all for your replies concerning measuring the refractive
index of a crystalline powder using a refractometer. The replies
indicated that it is not easy and has very specific requirement that I
can not obtain. So I will use a set of standard refractive index
liquids and a polarized light microscope which is the way I had
originally planned to go.

Thanks again for the comments, you folks are a veritable deep well of
knowledge.

Damian Neuberger




From: Brian G. Demczyk :      demczyk-at-ERXINDY.rl.plh.af.mil
Date: Fri, 14 Mar 1997 11:01:08 +0001 (EST)
Subject: Snappy Framegrabber board

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Some time back, the Snappy framegrabber board, from Play, Inc.
was recommended as an economical way to obtain digitized EM images.
Well, acting on this recommendation, I purchased a Snappy board. However,
I have been unable to obtain an acceptable stored image file, using
the Snappy board (and it's associated software) in conjunction
with a CCD camera and macro lens mounted above a light box on which the
EM negative is mounted. Previously, this technique has worked well,
when using another grabber board (costing around five times as much!)
and NIH Image software. Has anyone obtained reasonable captured images of
this type using the Snappy board? I fear that this application is beyond
the resolution capabilities of the Snappy board, but there is always the
possibility that I'm not setting something properly in the Snappy capture
software.




From: kennedy-at-nsi.edu (Grace Kennedy)
Date: Fri, 14 Mar 1997 08:29:25 -0800
Subject: uranyl acetate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Over the years I've found that uranyl acetate (dry powder/crystals) seem to
degrade with time and find the staining greatly attenuated. It becomes
less soluble and keeps for a shorter time in solution. Does anyone know
the mechanism(s) of this change and how to perhaps prevent it? I do store
dark, and have had it happen in both plastic and glass containers. I make
my staining solutions up in small volumes. I've ended up having more
picked up by EH&S than I've ever really used. If it's inevitable, maybe we
could ask our vendors to offer smaller quantities as an
alternative......Grace Kennedy






From: lporter-at-goodyear.com (LE Porter)
Date: Fri, 14 Mar 1997 12:24:44 -0500
Subject: Microscopy position available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy position available.

Please do not reply via eMail but direct your resume to:
M R Jobe Salaried Personnel
THE GOODYEAR TIRE & RUBBER COMPANY
1144 E. Market Street
Akron, Ohio 44316

The Goodyear Tire & Rubber Company, a world leader in tire development and
manufacturing, has an excellent opportunity for a professional skilled in
Microscopy in its Analytical Sciences Department, Corporate Research
Division, located in Akron, Ohio. This position involves the analysis and
characterization of polymers and materials used in the tire and rubber
industry.

Qualified candidates must have a PhD or MS with 3 years experience in
light, electron and atomic force microscopy with a strong background in
image analysis. Current experience should include techniques employed in
polarized light, brightfield, phase contrast, fluorescence, interference
microscopy and photomicrography. Knowledge of transmission and scanning
electron microscopy and EDS systems is also required. Familiarity with
failure analysis and mixing studies of rubber, plastics and composites is
desirable.

This position with a Fortune 100 industry leader offers excellent benefits,
relocation assistance and competitive salary commensurate with education
and experience. If you have the qualifications and desire to meet the
rewarding challenges presented by Goodyear, direct your resume to:

M R Jobe
Salaried Personnel
THE GOODYEAR TIRE & RUBBER COMPANY
1144 E. Market Street
Akron, Ohio 44316

An Equal Opportunity Employer, M/F/D/V
Applicants must be lawfully authorized to work in the U.S.

L E Porter Phone (330) 796-1620
Head of Microscopy Fax (330) 796-3304
The Goodyear Tire & Rubber Company EMail LPORTER-at-GOODYEAR.COM
Dept 415A
142 Goodyear Blvd
Akron, OH 44305
USA







From: Damian_Neuberger_at_RLT013-at-ccmailgw.mcgawpark.baxter.com
Date: Tue, 11 Mar 1997 14:49:52 -0600
Subject: Refractive Index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Good Afternoon (Central Std Time USA) all Microscopists:

I have just had a request to determine the refractive index of zinc
pyrithione, a crystalline, colorless, transparent powder. Does anyone
know what the RI is? It's not in McCrone Atlas. I will be getting a
sample Monday and with a set of Cargille RI liquids, I will determine
the RI.

Alternatively, I have a B & L Abbe type refractometer used to
determine refractive index of liquids. However, I thought I read
somewhere that it could be used to determine the RI of solids. If
this is in fact possible, does anyone out there know how?

TIA

Damian Neuberger
neuberd-at-baxter.com
#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: James Ekstrom :      jekstrom-at-exeter.edu (by way of Nestor J. Zaluzec)
Date: Tue, 11 Mar 1997 18:01:04 -0500
Subject: ISI SEM Manual needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

My wife is having to do a poll of scientist for one of her certification
classes. She has taught science in middle and high school. For the
philosophers in the crowd I hope that you can have fun with this one, if you
could help out with her survey it would be much appreciated. For all
others, I ask your indulgence. Please forward you replys to me and I will
deliver them. The survey follows.

Thank you,
Chuck


SURVEY QUESTIONS

Name:
Degree/Profession:

1. How would you describe the Nature of Science?

2. Describe the ideal classroom and curriculum for scientific learning.

3. What role do you believe the History of Science should play in teaching
science in the classroom?

END SURVEY
-------------------------------------
Name: Charles Gilbert VOC:(704)355-5261
Carolinas Medical Center FAX:(704)355-7996
Dept of Pediatric Research
PO Box 32861
Charlotte, NC 28232-2861


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I am acquiring an ISI SX-30E and need a copy of the manual for this
instrument without paying the hundreds of dollars Topcon wants for such
a manual.
Any suggestions where I might be able to get a xerox of the manual or
borrow a copy to xerox.
Any help would be appreciated.

Jim Ekstrom
Phillips Exeter Academy


#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: Sandra F. Zane :      sfzane-at-unccvm.uncc.edu
Date: Fri, 14 Mar 1997 14:28:01 -0600
Subject: Projection slide film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello Fellow Listserv Members,
For the past 15 or so years I have been using a wonderful film by
Kodak which makes very nice black and white slides of continous tone and
high contrast copy. This is a positive film which requires no reversal
processing. It is called Kodak Direct Duplicating Microfilm 2468. I have
been buying it from a supplier in Florida who has apparently gone out of
business. Now here is the catch. Kodak, according to my long time
photography needs supplier, is not willing to sell this film in anything
less than a case. There are 50 100ft rolls in a case at about $40.00 per
100ft roll. My question is....are there any of you out there who might be
familiar with this film and know of suppliers other than Brandon's in
Jacksonville, FLA.? Brandon's is the supplier from whom I have ordered this
film in the past. Or another question which has just come to mind...are
there any EM Suppliers out there who might be interested in this film as one
of their catalogue items? I would appreciate hearing from any of you who
might have information on where to find this film or someone who might be
interested in purchasing a case for distribution.
I trust this is an appropriate post. Since we all make slides from
time to time, I felt that it might be. Perhaps it would be better to reply
to me directly and if there are others interested, let me know and I will
pass on any information I receive. Thank you very much, Sandra Zane
Sandra F. Zane, EM Tech. sfzane-at-email.uncc.edu
Dept. of Biology, UNCC Ph.(704)547-4051
9201 University City Blvd. Fax (704)547-3128
Charlotte, NC 28223





From: bozzola-at-siu.edu (John J. Bozzola)
Date: Fri, 14 Mar 1997 15:03:23 -0600
Subject: Scanning TEM Negatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

We just received a Polaroid Sprintscan 45 for digitizing negatives from our
microscopes. Preliminary trials look good. Unfortunately, no negative
holder is available for TEM negatives (3 1/4 x 4 inch). This is really
unfortunate since we have hundreds of such negs to scan and using the 4x5
holder isn't very user friendly - although it can be done. Has anyone had
experience with this system? Any workarounds?

I suggested to Polaroid that if they would provide me with a couple of
their standard 2 1/4 x 2 3/4 inch adapters, that I could have our shop
mill out the proper size frame. I was shocked to find that not only were no
plans in place to make the TEM holders but that Polaroid would not provide
even the stock holders ( or even the metal) for modification. Caveat
emptor.





####################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Neckers Building, Room 146 - B Wing
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################






From: kszaruba-at-MMM.COM (Karen S. Zaruba)
Date: Fri, 14 Mar 1997 15:11:04 -0600
Subject: Re: Snappy Framegrabber board

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

About 6 months ago I asked the group a similar question about the Snappy and
got a number of good responses concerning quality of cable, sources of
noise, etc.

I have sent a rather long summary of all these replies to the originator of
this post, and just wanted to make everyone else aware it is available.
Just E-mail me and I'll forward it on to you individually (didn't want to
bother the whole listserve with such a long message).

-Karen


On March 14, 1997, demczyk-at-erxindy.rl.plh.af.mil wrote:

} Some time back, the Snappy framegrabber board, from Play, Inc.
} was recommended as an economical way to obtain digitized EM images.
} Well, acting on this recommendation, I purchased a Snappy board. However,
} I have been unable to obtain an acceptable stored image file, using
} the Snappy board (and it's associated software) in conjunction
} with a CCD camera and macro lens mounted above a light box on which the
} EM negative is mounted. Previously, this technique has worked well,
} when using another grabber board (costing around five times as much!)
} and NIH Image software. Has anyone obtained reasonable captured images of
} this type using the Snappy board? I fear that this application is beyond
} the resolution capabilities of the Snappy board, but there is always the
} possibility that I'm not setting something properly in the Snappy capture
} software.
}
}
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Life Sciences Sector Lab Reply: kszaruba-at-mmm.com
3M Company
3M Center 270-1S-01 Phone: 612-737-2971
St. Paul, MN 55144-1000 Fax: 736-1519

These opinions are my own and may not represent those of 3M.







From: WARRENJ1-at-cliffy.polaroid.com
Date: 3.14.97 4:03 PM
Subject: Scanning TEM Negatives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John

I am sorry that you were given the information that you were given. I
recommended earlier this year that we offer a carrier for TEM
negatives due to the lack of other currently available options. The
last word I heard was that we are working on a carrier that will be
adjustable for various film sizes. They also are working on creating a
carrier dedicated to the TEM format you mentioned. As a temporary
solution, cutting a mask that would fit in the 4x5 holder with the
proper opening should work.

I am going to forward your message to the Worldwide Product &
Technical Managers for Polaroid Scanners.

I'll keep you posted.

John D. Warren
Area Sales Manager
Digital Products
Polaroid Corporation


______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

We just received a Polaroid Sprintscan 45 for digitizing negatives from our
microscopes. Preliminary trials look good. Unfortunately, no negative
holder is available for TEM negatives (3 1/4 x 4 inch). This is really
unfortunate since we have hundreds of such negs to scan and using the 4x5
holder isn't very user friendly - although it can be done. Has anyone had
experience with this system? Any workarounds?

I suggested to Polaroid that if they would provide me with a couple of
their standard 2 1/4 x 2 3/4 inch adapters, that I could have our shop
mill out the proper size frame. I was shocked to find that not only were no
plans in place to make the TEM holders but that Polaroid would not provide
even the stock holders ( or even the metal) for modification. Caveat
emptor.





####################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Neckers Building, Room 146 - B Wing
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################




From: WARRENJ1-at-cliffy.polaroid.com
Date: 3.14.97 3:28 PM
Subject: Projection slide film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Have you tried Polaroid's 35mm Polagraph Instant Slide Film? It is a
high contrast b&w film that you can process & mount at your desk. The
processing does require our processor - either power or manual. The
film comes with the development chemistry. You just have to buy the
reusable slide mounts separately. It is a little denser than wet
processed film, but it may not be as apparent with a high contrast
image.

John Warren
Area Sales Manager
Digital Products
Polaroid Corporation


______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hello Fellow Listserv Members,
For the past 15 or so years I have been using a wonderful film by
Kodak which makes very nice black and white slides of continous tone and high
contrast copy. This is a positive film which requires no reversal
processing. It is called Kodak Direct Duplicating Microfilm 2468. I have
been buying it from a supplier in Florida who has apparently gone out of
business. Now here is the catch. Kodak, according to my long time
photography needs supplier, is not willing to sell this film in anything less
than a case. There are 50 100ft rolls in a case at about $40.00 per 100ft
roll. My question is....are there any of you out there who might be familiar
with this film and know of suppliers other than Brandon's in Jacksonville,
FLA.? Brandon's is the supplier from whom I have ordered this film in the
past. Or another question which has just come to mind...are there any EM
Suppliers out there who might be interested in this film as one of their
catalogue items? I would appreciate hearing from any of you who might have
information on where to find this film or someone who might be interested in
purchasing a case for distribution.
I trust this is an appropriate post. Since we all make slides from
time to time, I felt that it might be. Perhaps it would be better to reply
to me directly and if there are others interested, let me know and I will
pass on any information I receive. Thank you very much, Sandra Zane
Sandra F. Zane, EM Tech. sfzane-at-email.uncc.edu
Dept. of Biology, UNCC Ph.(704)547-4051
9201 University City Blvd. Fax (704)547-3128
Charlotte, NC 28223




From: richard.easingwood-at-stonebow.otago.ac.nz (Richard Easingwood)
Date: Wed, 12 Mar 1997 17:04:19 +1200
Subject: Vascular casting - Mercox supplier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: st004718-at-brandywine.otago.ac.nz
Message-Id: {v01540b05af4be81f8e45-at-[139.80.34.57]}
Mime-Version: 1.0

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Does anyone who performs vascular casting know where to obtain a product
called Mercox CL2? A fellow electron microscopist has heard that it is the
latest thing in vascular casting when mixed with methyl methacrylate but
hasn't been able to get any further info.
If anyone knows who makes it and where it can be bought we would like to hear.

Thanks in advance

Richard Easingwood
South Campus Electron Microscope Unit
School of Medical Sciences
University of Otago
PO Box 913
Dunedin
NEW ZEALAND

Telephone: 64-03-479 7301
Facsimile: 64-03-479 7254

e-mail: richard.easingwood-at-stonebow.otago.ac.nz




#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: Gary Dietrich Chinga :      garyc-at-stud.ntnu.no
Date: Wed, 12 Mar 1997 18:31:10 +0100 (MET)
Subject: Tonicity and osmolarity?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hi!

Can someone explain the relation between the tonicity and osmolarity
of a fixation solution?


Gary.


#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Tue, 11 Mar 1997 14:48:36 -0800 (PST)
Subject: CPD Thanks!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Thanks to all who responded to my cpd inquiry. The consensus seems to be
that fresh, dry 100% ETOH is very important and that I need to keep my
samples longer in the liquid CO2 to get a better exchange with the ETOH.

I'll try out all the suggestions. Hopefully, the only raisins I'll get
from now on will be in my breakfast cereal. = )


Thanks again.


Paula = )


p.s. I'm going to phone some vendors of the resins & such to see if
they've done studies as to which type of gloves work best. I'll keep you
posted.

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu


#############################
Notice:

This message was found in a dead-letter box and appears to be for you.
If you have already gotten a copy of this message, we beg your tolerance.

The Unix Systems Group





From: RA :      ralpha-at-softcom.net
Date: Sun, 16 Mar 1997 16:03:37 -0800
Subject: Wang model 3007 Microscopes.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Does anyone have any experience with this microscope. Is this
considered a fine microscope for the microscopy hobby? Is this scope
lacking any of the essentials or necessities for the amateur
microscopist? I was surfing the net under microscopes and this one really
caught my attention. Any comments would be greatly appreciated.

Thanks

Ralph




From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Sun, 16 Mar 1997 21:16:01 -0400
Subject: unstable ion pumps

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

A few days ago someone asked about having sputter ion pumps rebuilt. It
occurs to me to comment that sometimes it is possible to squeeze a few
months of extra use out of a pump before rebuilding becomes absolutely
necessary. Frequently, pumps become unstable because a low impedence path
develops between the anode and cathode structures, usually because a flake
of Ti breaks off the anode and lodges between it and the cathode, or
because of the growth of a Ti whisker between the two electrodes.
Sometimes such a condition can be relieved temporairly by turning the
power to the pump on for a few seconds (ONLY a FEW) when the pressure in
the pump is just above 1 Pa (0.01 Torr), repeating the process three or
four times, if necessary. (See Vac. Methods in EM, p 295) If this
treatment is successful, the pump's performance will stabalize, and the
pump will perform satisfactorily again, but only for a limited time -
ultimately, of course, it will need to be reconditioned.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: xia chen :      xiachen-at-mail.med.cornell.edu
Date: Mon, 17 Mar 1997 08:46:48 -0500
Subject: Help NIH Image

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi,
Help! I just started to use the NIH-image to analyze the distribution of
gold particles on the EM negatives. The software was downloaded from the
web site and it is NIH 1.61. My problem at this point is that I can not
open my scanned image to the proper size I wanted. I tried to acquire, open
or import the image, the result is the same -- a very small picture show up
on the screen. When this image was magnified, the pixels came out which
make particle counting impossible. Apparently, it is not the problem with
the scanner since the image could be processed nicely with Adobe. But if I
open the Adobe processed image through NIH, the same thing happens as
described above. I also tried to scan the negatives using higher ppi,
approximately 1000 ppi to incease the resolution. That did not help.
Any suggestions and help will be greatly appreciated.

Xia Chen
Dept. of Pathology
CUMC, N.Y.






From: pat_masarachia-at-merck.com (Pat Masarachia)
Date: Mon, 17 Mar 1997 10:01:24 -0500 (EST)
Subject: dirt on LR White sections with Immuno gold labeling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

When I label LR White thin sections of decalcified bone with
immunogold I get a fine contamination - dirt - that is not present on
my conventional EM/epon sections.
The contaminant is a fine floccular stuff made up of filament-
like strands, each less than 10 nm in diameter, that clump togethr
into irregular shapes of widely ranging size. The contaminant is not
seen on the phosphorescent viewing screen but definitely shows up
in the negative and print.
All diluents and wash solutions are filtered with 0.2 micron
Gelman Acrodisc syringe filters. Antibodies are diluted either in Tris-
saline or BioMeda Primary Antibody Diluent; blockers are serum, fish
gelatin, BSA. Wash solutons ar tris-saline. All solutions have
Triton X. Final staining is with methanolic or aqueous uranyl acetate.
Deleting uranyl acetate staining still leaves contamination.
Deleting primary antibody still leaves contamination. LR White alone
without processing for immuno and stained with uranyl acetate is clean.
So, something in the gold, or in the secondary antibody may be
doing it, but I haven't checked this out yet. Any ideas on what this
dirt could be and how to get rid of it would be appreciated.
Sorry for the long-windedness...and Thanks.

Pat Masarachia
Bone Biology
Merck Research Labs
West Point, PA
tel:215-652-7999
e-mail: pat_masarachia-at-merck.com






From: rnessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Mon, 17 Mar 1997 12:14:16 -0500 (CDT)
Subject: Immunocytochemistry Workshop to be held

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The Central Microscopy Research Facility at the University of Iowa
will host an immunocytochemistry workshop given by Jan Leunissen and
Peter Van der Blass. It is to be held May 9-10, 1997 in Iowa City.
For more information, visit our website at:
http://www.uiowa.edu/~cemrf/cemrf/jan_workshop.html
or contact Kenneth Moore (319)-335-8143, kenneth-moore-at-uiowa.edu

Randy Nessler rnessler-at-emiris.iaf.uiowa.edu
Views expressed are my own.





From: Ya Chen :      ychen14-at-facstaff.wisc.edu
Date: Mon, 17 Mar 1997 11:53:05 -0600
Subject: Re: Help NIH Image

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

NIH Image is extensively used in the Integrated Microscopy Resource (IMR).
We never have this kind of problem. If you want, I can FTP you some of my
images. In this way you can easily find where the problem is. Hope that
will help you.

Ya Chen


Ya Chen

=========================================================================
\ / Integrated Microscopy Resource (IMR)--
\ / __ an NIH Biomedical Research Resource TEL : 608-263-8481
\/ / / University of Wisconsin-Madison FAX : 608-265-4076
/ / / 1675 Observatory Drive #159 Email1:ychen14-at-facstaff.wisc.edu
/ /__/_ Madison, WI 53706 Email2:chen-at-calshp.cals.wisc.edu
=========================================================================
IMR WWW Home Page: http://www.bocklabs.wisc.edu/imr.html






From: nyao-at-Princeton.EDU (Nan Yao)
Date: Mon, 17 Mar 1997 16:11:41 -0500 (EST)
Subject: Wanted: holders and parts for the LEO/Zeiss 910 TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Does anyone have available (new or used) a double tilt holder, a double
tilt heating holder, a double tilt cooling holder and a selective aperture
set for the LEO/Zeiss 910 TEM? If so, please let me know (incl. price).

Thank you.

Nan Yao






From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Mon, 17 Mar 1997 16:33:06 -0800
Subject: re: positive film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear List:

Sandra Zane was wondering where she could get Kodak Direct
Duplicating Microfilm 2468. We use Eastman 5360 which sounds like the
same thing; 35 mm, gives a positive without a complicated kit, developes
in Dektol. We order it locally (I think) or you can get it from Freestyle
Sales Co. 5124 Sunset Blvd. Los Angeles, CA 90027,
www.freestylesalesco.com. In their catalogue it is called "Kodak B&W
duplicating film 5360" and is available in 50 or 100 foot rolls or 4x5
sheets.

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (908)-235-4583; fax -4029 e-mail: mcauliff-at-umdnj.edu
***************************************************************




From: Steven Lockett :      lockett-at-white.lbl.gov
Date: Mon, 17 Mar 1997 17:05:26 -0800
Subject: postdoc opening in image analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

POSTDOCTORAL POSITION is available immediately to develop three
dimensional image analysis algorithms for quantitative analysis
of cells in tumor specimens. Experience in image analysis and UNIX
workstations, and a strong math background are essential. Candidate
should have a recent Ph.D. in the physical sciences, or
Computer Science and should be an expert in developing code in
C/C++. Send curriculum vitae and the names of three referees to:

Dr. Ravi Malladi,
MS 50A-2152
Lawrence Berkeley National Laboratory
University of California,
Berkeley, CA 94720.






From: Carlos E. Barbosa :      grial-at-satlink.com
Date: Mon, 17 Mar 1997 22:31:16 -0300
Subject: LM: Info on cathodoluminiscence

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear netters!

Does anyone knows about providers of cathodoluminiscence equipments for
LM.

Thank you in advance!

Carlos E. Barbosa
Email: grial-at-satlink.com





From: RA :      ralpha-at-softcom.net
Date: Mon, 17 Mar 1997 19:57:38 -0800
Subject: Plan Vs Semi-plan objectives

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Forgive my inexperiece and knowledge on this new hobby of mine but can
someone briefly explain the differece between semi-plan objectives and
plan objectives. Under what applications would one type be more
advantageous over the other type.
Thanks for any information you may have on this subject.

Ralph




From: Peggy Brannigan :      brannign-at-asrr.arsusda.gov
Date: Tue, 18 Mar 1997 09:39:48 -0500
Subject: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello immunogold experts!

Hope I'm not duplicating an earlier topic but couldn't find anything in
postings kept over the last few months so, here goes...

What factor (s) in an immungold labeling protocol has (have) the greatest
impact on background, e.g.. antibody dilution, buffer, etc.? I'm
looking at the immunogold protocol itself so I can use valuable tissue
already embedded.

(I'm labeling virus-infected plant leaf tissue with monoclonal antibodies
to viral proteins - samples are embedded in LR White and fixed in 3%
paraformaldehyde/0.2% glutaraldehyde. We block before primary antibody
incubation with a TRIS-HCl buffer containing bovine serum albumin and goat
normal serum. Our secondary antibody is a commercially prepared goat
anti-mouse gold conjugate. Usually we have excessive background on
chloroplasts, mitochondria and cell walls; if we get rid of background by
increasing dilutions of antibodies, we lose specific reactions as well).


One interesting point....we don't have this problem with polyclonal
antibodies, only monoclonals.

Thanks for any tips -- this is really frustrating!


Peggy

Peggy Brannigan
Electron Microscopy
Floral and Nursery Plants Research Unit
National Arboretum

Bldg. 010A R.238
10300 Baltimore Avenue
Beltsville, MD USA20705

Phone: (301) 504-6097
Fax : (301) 504-5096
Email: brannign-at-asrr.arsusda. gov







From: Sara Miller :      saram-at-acpub.duke.edu
Date: Tue, 18 Mar 1997 10:06:57 -0500 (EST)
Subject: re: positive film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Mon, 17 Mar 1997, Geoff McAuliffe wrote:

} Date: Mon, 17 Mar 1997 16:33:06 -0800
} From: Geoff McAuliffe {mcauliff-at-UMDNJ.EDU}
} To: MSA Listserver {Microscopy-at-Sparc5.Microscopy.Com}
} Subject: re: positive film
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Dear List:
}
} Sandra Zane was wondering where she could get Kodak Direct
} Duplicating Microfilm 2468. We use Eastman 5360 which sounds like the
0} same thing; 35 mm, gives a positive without a complicated kit, developes
} in Dektol. We order it locally (I think) or you can get it from Freestyle
} Sales Co. 5124 Sunset Blvd. Los Angeles, CA 90027,
} www.freestylesalesco.com. In their catalogue it is called "Kodak B&W
} duplicating film 5360" and is available in 50 or 100 foot rolls or 4x5
} sheets.
}
} Geoff
} --
} ***************************************************************
} Geoff McAuliffe, Ph.D.
} Neuroscience and Cell Biology
} Robert Wood Johnson Medical School
} 675 Hoes Lane Piscataway, NJ 08854
} voice: (908)-235-4583; fax -4029 e-mail: mcauliff-at-umdnj.edu
} ***************************************************************
} We have good success with 5360 also, and it's easy to get. Here is a
copy of information I sent directly to Sandra:

The Direct MP Film 5360 is cat # 24611 from Ted Pella, $22.90/100 ft
roll. 800 237-3526 or tedpel-at-aql.com or Fax 916 243-3761 I put the
camera on B and manually work the shutter for 2, 2.5, 3 sec, depending on
the size (height of camera). It's so cheap, I usually bracket and shoot 3
shots for each to make sure I don't have to repeat shooting. Most of the
time, any of the shots could be used, but I pick the ones that are most
closely matched for use in a single presentation. I develop in D-19 4
min at 20 oC. There is another film, Kodak Contrast Copy Film, that works
about the same, but gives a bluer tint. I only keep the 5360 now.

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735





From: Robert Underwood :      underwoo-at-u.washington.edu
Date: Tue, 18 Mar 1997 08:37:23 -0800 (PST)
Subject: Re: dirt on LR White sections with Immuno gold labeling

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Pat,

I have done a lot of LR White immunostaining and have had many kinds of
dirt problems. It sounds like you are doing nearly everything right but
you didn't mention doing a final rinse in filtered distilled water after
your gold secondary. I rinse 2-3x in my buffer after secondary then do a
fairly vigorous rinse in the beaker of distilled filtered water, then dry
the grids, then go to staining.

You may also check your water and photoflo source and see if it is
actually on your negatives.

Bob
Morphology Core
U.of Washington
Seattle

On Mon, 17 Mar 1997, Pat Masarachia wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} When I label LR White thin sections of decalcified bone with
} immunogold I get a fine contamination - dirt - that is not present on
} my conventional EM/epon sections.
} The contaminant is a fine floccular stuff made up of filament-
} like strands, each less than 10 nm in diameter, that clump togethr
} into irregular shapes of widely ranging size. The contaminant is not
} seen on the phosphorescent viewing screen but definitely shows up
} in the negative and print.
} All diluents and wash solutions are filtered with 0.2 micron
} Gelman Acrodisc syringe filters. Antibodies are diluted either in Tris-
} saline or BioMeda Primary Antibody Diluent; blockers are serum, fish
} gelatin, BSA. Wash solutons ar tris-saline. All solutions have
} Triton X. Final staining is with methanolic or aqueous uranyl acetate.
} Deleting uranyl acetate staining still leaves contamination.
} Deleting primary antibody still leaves contamination. LR White alone
} without processing for immuno and stained with uranyl acetate is clean.
} So, something in the gold, or in the secondary antibody may be
} doing it, but I haven't checked this out yet. Any ideas on what this
} dirt could be and how to get rid of it would be appreciated.
} Sorry for the long-windedness...and Thanks.
}
} Pat Masarachia
} Bone Biology
} Merck Research Labs
} West Point, PA
} tel:215-652-7999
} e-mail: pat_masarachia-at-merck.com
}
}
}





From: Robert Underwood :      underwoo-at-u.washington.edu
Date: Tue, 18 Mar 1997 08:50:13 -0800 (PST)
Subject: Re: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Peggy

I would think if your minus primary control are clean then your primary
antibody is non-specificly binding to something in the plant tissue.
Usually the primary has the most effect on backround. Is your antisera
purified?

Bob
Morphology Core
U of Washington
Seattle

On Tue, 18 Mar 1997, Peggy Brannigan wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello immunogold experts!
}
} Hope I'm not duplicating an earlier topic but couldn't find anything in
} postings kept over the last few months so, here goes...
}
} What factor (s) in an immungold labeling protocol has (have) the greatest
} impact on background, e.g.. antibody dilution, buffer, etc.? I'm
} looking at the immunogold protocol itself so I can use valuable tissue
} already embedded.
}
} (I'm labeling virus-infected plant leaf tissue with monoclonal antibodies
} to viral proteins - samples are embedded in LR White and fixed in 3%
} paraformaldehyde/0.2% glutaraldehyde. We block before primary antibody
} incubation with a TRIS-HCl buffer containing bovine serum albumin and goat
} normal serum. Our secondary antibody is a commercially prepared goat
} anti-mouse gold conjugate. Usually we have excessive background on
} chloroplasts, mitochondria and cell walls; if we get rid of background by
} increasing dilutions of antibodies, we lose specific reactions as well).
}
}
} One interesting point....we don't have this problem with polyclonal
} antibodies, only monoclonals.
}
} Thanks for any tips -- this is really frustrating!
}
}
} Peggy
}
} Peggy Brannigan
} Electron Microscopy
} Floral and Nursery Plants Research Unit
} National Arboretum
}
} Bldg. 010A R.238
} 10300 Baltimore Avenue
} Beltsville, MD USA20705
}
} Phone: (301) 504-6097
} Fax : (301) 504-5096
} Email: brannign-at-asrr.arsusda. gov
}
}
}
}





From: GREG VAUGHN MARTIN :      gmartin-at-welchlink.welch.jhu.edu
Date: Tue, 18 Mar 1997 11:50:43 -0500 (EST)
Subject: Re: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hey Peggy --

We also find that immuno-EM is problematic with MAbs. Are you
using ascites or an IgG prep? I would use IgG if possible (the EZ Sep
products from Pharmacia Biotech are handy for isolating IgG from
ascites, MHO). You may want to try adding 0.05% Tween-20 to your blocking
and antibody soln's. This often helps bring down the background. The
downside (there's always a downside) is that the tween also tends to wash
a lot of the contrast out of the sections -- at least it does for our
LRGold embedded material. You may need to adjust the length of time you let
the grids float on the antibody sol'n to give good label without too much loss
of contrast. 1-4 hours on the primary is a good range for starters.

Hope this helps,

Greg Martin
Dept. of Cell Biology and Anatomy
Johns Hopkins School of Medicine


On Tue, 18 Mar 1997, Peggy Brannigan wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello immunogold experts!
}
} Hope I'm not duplicating an earlier topic but couldn't find anything in
} postings kept over the last few months so, here goes...
}
} What factor (s) in an immungold labeling protocol has (have) the greatest
} impact on background, e.g.. antibody dilution, buffer, etc.? I'm
} looking at the immunogold protocol itself so I can use valuable tissue
} already embedded.
}
} (I'm labeling virus-infected plant leaf tissue with monoclonal antibodies
} to viral proteins - samples are embedded in LR White and fixed in 3%
} paraformaldehyde/0.2% glutaraldehyde. We block before primary antibody
} incubation with a TRIS-HCl buffer containing bovine serum albumin and goat
} normal serum. Our secondary antibody is a commercially prepared goat
} anti-mouse gold conjugate. Usually we have excessive background on
} chloroplasts, mitochondria and cell walls; if we get rid of background by
} increasing dilutions of antibodies, we lose specific reactions as well).
}
}
} One interesting point....we don't have this problem with polyclonal
} antibodies, only monoclonals.
}
} Thanks for any tips -- this is really frustrating!
}
}
} Peggy
}
} Peggy Brannigan
} Electron Microscopy
} Floral and Nursery Plants Research Unit
} National Arboretum
}
} Bldg. 010A R.238
} 10300 Baltimore Avenue
} Beltsville, MD USA20705
}
} Phone: (301) 504-6097
} Fax : (301) 504-5096
} Email: brannign-at-asrr.arsusda. gov
}
}
}
}




From: Alwyn Eades :      eades-at-uimrl7.mrl.uiuc.edu
Date: Tue, 18 Mar 1997 11:32:38 -0600
Subject: Job Vacancy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Microscopy and Computing

I am looking for a person to work on an exciting new project.

The appointment could be at the post-doctoral level or at other levels
according to the background of the person appointed. In any case the post
will be for three years.

The job is at the Materials Research Laboratory of the University of
Illinois at Urbana.

The project is a joint enterprise involving Argonne National Lab, Oak Ridge
National Lab, the Lawrence Berkeley Lab and NIST as well as the University
of Illinois. The Project has the aim of developing a new kind of
environment for electron microscopy and related techniques, in which the
instruments can be operated remotely with the same effectiveness as they can
be operated in the instrument room. More details of the project can be found
at http://tpm.amc.anl.gov/MMC MMC is the abbreviation of the project name.

I am looking for someone who has familiarity with electron microscopy
(preferably TEM) or a closely related technique - and who has well developed
interests and experience in computing, particularly the interfacing of
instruments for computer control and/or the networking of images.

Will any one interested please contact me right away. We would like the job
to be started as soon as possible.
**
Alwyn Eades Center for Microanalysis of Materials
University of Illinois at Urbana-Champaign
Phone 217 333 8396 Fax 217 244 2278
eades-at-uimrl7.mrl.uiuc.edu (NB those are letter l not ones)
**





From: Arizona Imaging and Microanalysis Society :      dcromey-at-ccit.arizona.edu
Date: Tue, 18 Mar 1997 11:18:18 -0600
Subject: AIMS mtg schedule 3/27-3/28/97

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

You are invited to attend the 1997 Annual Spring meeting of the Arizona
Imaging and Microanalysis Society. The meeting will be Thursday March 27th
and Friday March 28, 1997. Meetings will be held in the Senior Ballroom of
the University of Arizona's Student Union Building. Admission is free to
interested participants.

Please pass this information on to interested colleagues.

=========================================================================
DIGITAL IMAGING
=========================================================================

Digital imaging has found its way into many scientific fields (biology,
engineering, astronomy, radiology, etc.). As a wider range of researchers
begin to use these techniques, they are often unsure of where to begin and
are unaware of the pitfalls involved in the field of digital imaging. This
workshop will introduce you to some important concepts and issues relevant
to digital imaging.

Thursday, March 27, 1997

9:00 Welcome and overview of the meeting
Doug Cromey, President, AIMS

9:15 Image acquisition: General approaches, common problems and issues
relating to CCD cameras.
Scott Sternberg, Photometrics

10:15 Coffee break

10:30 Data output devices
John Spenseri, Micrographix West

11:15 Data storage, movement and archiving
Doug Cromey, President, AIMS

12:00 Lunch (on your own)

1:30 Image file formats, cross-platform issues, 3-D reconstruction
Marvin Landis, CCIT-User Support

2:30 Poster session, Vendor presentations

3:15 Coffee Break

3:30 Poster session, Vendor presentations

6:00 Buffet dinner in the Student Union's Sr. Ballroom
(Mexican-style entrees, Preregister with Patty Jansma, cost $11.00)


Friday, March 28, 1997

9:00 Ethics and image analysis: Panel discussion.
(Questions from the audience are welcome.)

Panel members:
John Gilkey, UA, Pharmacology
Marvin Landis, UA, CCIT- User Support
David Ring, Dir. Photography, Biomedical Communications, UA
Mary Rykowski, Assist. Prof., Cell Biology and Anatomy, UA
Scott Sternberg, Photometrics

10:30 Coffee Break

10:45 Student platform presentations

11:45 Lunch. AIMS Business Meeting
New officers installation.

1:00 Student platform presentations

3:00 Presentation of student awards and closing remarks


Contact Information:
Doug Cromey AIMS President (520)626-2824
Patty Jansma AIMS President-elect (520)621-6671
Lynne Oland AIMS Secretary (520)621-6671
===============================================================
ARIZONA IMAGING & MICROANALYSIS SOCIETY
WWW: http://www.msa.microscopy.com/MSALAS/AIMS/AIMSHOME.HTM
email: aims-at-msa.micrscopy.com





From: hall-at-aecom.yu.edu (Dr. David Hall)
Date: Tue, 18 Mar 1997 13:08:18 -0500 (EST)
Subject: histochemical stains for plastic sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

A histochemistry query: In the course of our work, we examine many
different vertebrate tissues (brain, skin, liver, kidney, etc) embedded in
plastic resin (Eponate or LR White). We currently do a lot of LM viewing of
2 micron sections, using toluidine blue as a counterstain. Some tissues
have been fixed with aldehydes and osmium, for subsequent EM thin section
studies. For immunocytochemistry, other tissues are only lightly fixed with
aldehydes.

Our question:
Can you suggest other useful counterstains, besides toluidine blue,
that are compatible with plastic sections. If you have a favorite recipe,
please send to us off-line, so that we don't clog the listserver. We'll
compile a list of the answers we receive, for those who are interested.

Thanks in advance

David H. Hall hall-at-aecom.yu.edu
Dept of Neuroscience
Albert Einstein Col Medicine, Bronx, NY 10461
(718) 430-2195
(718) 430-8821 FAX





From: tnjed-at-uic.edu (Jack Gibbons)
Date: Tue, 18 Mar 1997 12:34:15 -0600
Subject: yeast morphology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



--============_-1353412441==_============
Content-Type: text/plain; charset="us-ascii"

I am doing a tem project with yeast and will have to identify membranes and
organells. Can anyone suggest a good book on yeast morphology?



--============_-1353412441==_============
Content-Type: application/mac-binhex40; name="In.toc"
Content-Disposition: attachment; filename="In.toc"

(This file must be converted with BinHex 4.0)



--============_-1353412441==_============
Content-Type: text/plain; charset="us-ascii"

Jack Gibbons
tnjed-at-uic.edu



--============_-1353412441==_============--





From: A Wilson :      awilson-at-aw.u-net.com
Date: Tue, 18 Mar 1997 18:49:55 +0000
Subject: 2nd RFD: sci.bio.immunocytochem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


BASIC BIOLOGICAL TRANSMISSION ELECTRON MICROSCOPY



A formal proposal to create a new group tentatively called
sci.bio.immunocytochem has just been posted to
news.announce.newgroups. This is a reposting of that proposal.

REQUEST FOR DISCUSSION (RFD)
unmoderated group sci.bio.immunocytochem

This is the 2nd Request For Discussion (RFD) for the creation of a
world-wide unmoderated Usenet newsgroup sci.bio.immunocytochem,
currently being discussed in news.groups

Suggestions for improvements to this proposal are welcome. Discussion
about it should take place in news.groups. A vote is expected to be
held in about three to four weeks.

This is not a Call for Votes (CFV); you cannot vote at this time.
Procedural details are below.

CHANGES from previous RFD:

2nd RFD posted because more than 60 days have elapsed since 1st RFD.
There are (minor changes in Distribution and Newsgroup line.

Newsgroup line:
sci.bio.immunocytochem Immuno-labelling of biological material.

RATIONALE: sci.bio.immunocytochem

Immunohistochemists and immunocytochemists already enjoy the benefits
of online communication, utilizing e-mail, accessing web sites, and
subscribing to specialised mailing lists. Usenet newsgroups are also
popular, but this is less obvious because articles with
immunocytochemical/immunohistochemical content get posted to many
different newsgroups. Most articles are posted to a favourite five or
six newsgroups including bionet.cellbiol, sci.med.immunology and
sci.techniques.microscopy, but often articles get posted to any one of
fourteen or fifteen newsgroups in the sci. and bionet. heirarchies.
Some of these are listed in the distribution list at the end of this
proposal.

In my view, no existing newsgroup fulfils the criteria necessary to
attract all the various immunocytochemistry postings. I do not wish
to draw users away from other newsgroups, only to encourage scientists
to share their knowledge and expertise on immunocytochemistry in the
most effective manner. In response to my proposal to create a
newsgroup dedicated to the discussion of immunocytochemistry and
immunohistochemistry, I have received e-mail and faxes from
researchers all over the world offering their support and
encouragment.

Immunocytochemistry and immunohistochemistry are not subdivisions of
immunology, molecular biology or chemistry. Microscopy, although
essential, is only a small part of the story. Immunocytochemistry and
immunohistochemistry are multi- disciplinary, therefore discussions
are destined to stay distributed amongst the different newsgroups
until they are all brought together under one umbrella. This would
then act as a focus point for all the immunocytochemists who are
already Internet users, and encourage new subscribers to Usenet.

CHARTER: sci.bio.immunocytochem

This is a newsgroup for the exchange of information relating to
immunocytochemistry and immunohistochemistry. This unique research
tool is used to locate and identify specific molecules in biological
material, at the microscopical level.

Articles posted to this group must be relevant to one or more aspects
of the above. The kind of subjects that may be discussed include
techniques, theory, presentation of results, requests for
collaboration, history, equipment, publication references, notice of
events, tips and trouble-shooting, jobs offered andwanted, jokes,
stories and new ideas, so long as the posting bears a direct relevance
to the central theme. There will be a list of Frequently Asked
Questions (FAQs) to help newcomers.

A relevant posting could just be a simple question or answer, for
example "Has anyone got any experience with this reagent ?"or "Which
course could I attend to learn more about immunogold labelling?".
There will be articles reminding people to read the list of FAQs prior
to posting their own article. Usenet readers may get involved in
complex discussions about, for example, multiple labelling, proper use
of control experiments, microwave antigen retrieval or quantitative
measurements. Remember that articles posted to a newsgroup are
intended for a wide readership, so if you have information which
concerns only one or two people then please don't use this newsgroup,
use e-mail.

Commercial advertisements for services, equipment or reagents violate
the charter unless one or more of the following apply: (a)The
advertisement is part of a comprehensive article designed specifically
to address issues raised in earlier articles posted to the group (b)A
general reference to the type of product does not suffice for
technical reasons and it is necessary to specify the exact commercial
product (c) The information is offered primarily for the benefit of
the readers (d)The advertisement is for second-hand equipment specific
to immunocytochemistry (e) Requests or offers for free products are
acceptable if they are not part of a sales promotion.

END CHARTER.

PROCEDURE:

This is a request for discussion, not a call for votes. In this phase
of the process, any potential problems with the proposed newsgroups
should be raised and resolved. The discussion period will continue
for a minimum of 21 days (starting from when the 2nd RFD for this
proposal is posted to news.announce.newgroups), after which a Call For
Votes (CFV) may be posted by a neutral vote taker if the discussion
warrants it. Please do not attempt to vote until this happens.

All discussion of this proposal should be posted to news.groups. This
RFD attempts to comply fully with the Usenet newsgroup creation
guidelines outlined in "How to Create a New Usenet Newsgroup" and "How
to Format and Submit a New Group Proposal". Please refer to these
documents (available in news.announce.newgroups) if you have any
questions about the process.

DISTRIBUTION:

This RFD has been posted to the following newsgroups:
news.announce.newgroups,news.groups,bionet.cellbiol,
bionet.diagnostics,bionet.immunology,bionet.microbiology,
bionet.molbio.methds-reagnts,sci.bio.misc,sci.med.immunology,
sci.techniques.microscopy

This RFD will be reposted to the following newsgroups after its
posting in news.announce.newgroups:
bionet.molbio.proteins,bionet.neuroscience,bionet.plants,
sci.bio.microbiology,sci.med,sci.med.laboratory,sci.misc,
sci.nanotech

This RFD will also be reposted to the following mailing lists after
its posting in news.announce.newgroups:

Histonet mailing list: {histonet-at-pathology.swmed.edu}
Information pertaining to the technical aspects of histology and
histopathology such as tissue fixatives and processing, routine
histology, special stains, immunohistochemistry, in-situ
hybridization etc. To subscribe type "subscribe digest" into the
subject box and leave the text box empty, or to subscribe to the full
service just type "subscribe". For more info access web site
http://www.mwrn.com/subject/histonet.htms

Microscopy Society of America listserver:
Questions/comments/answers in the various fields of Microscopy
Currently over 3000 subscribers. To subscribe send the message
"subscribe" to {Listserver-at-MSA.Microscopy.Com}
then send messages in plain text to
{Microscopy-at-MSA.Microscopy.Com}
For more info access web site http://www.amc.anl.gov/
Docs/anl/Nestor/Software/telecommList.html

Stanford University list server
To subscribe, send a message to
{majordomo-at-pathology.stanford.edu} with "subscribe ipox-l" in
the body of your message. This list helps pathologists and other
laboratory professionals to exchange information about
immunoperoxidase methods.

This RFD will also be reposted to the following web-sites after its
posting in news.announce.newgroups:

Royal Microscope Society
http://www.rms.org.uk
Web Master Dr R. A. D. Mackenzie
{r.a.mackenzie-at-open.ac.uk}

Center for Cell Imaging Department of Cell Biology
Yale University School of Medicine
Introduction to Immunocytochemistry
http://info.med.yale.edu/cellimg/CCIimmuno.html
Web Master Paul Webster
{ paul_webster-at-yale.edu}

Proponent: Amanda Wilson {awilson-at-aw.u-net.com}
Proponent: Paul Monaghan { monaghan-at-icr.ac.uk}
Mentor: Jonathan Grobe {grobe-at-netins.net}

Amanda Wilson
Deputy Manager, E.M. Unit
St George's Hospital Medical School,
S.W.London, UK
Tel: 0181 725 5220 (work)
e-mail {awilson-at-aw.u-net.com}





From: Scott Schwinge :      schwinge-at-fhl.washington.edu
Date: Tue, 18 Mar 97 12:57:30 -0800
Subject: Philips Model 300

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Greetings!

I am in need of an X-unit (EHT regulator) for our Philips Model 300. Does
anyone happen to know where I might find a working unit? Please respond
directly to me. Thanks in advance!

Scott Schwinge
University of Washington
Friday Harbor Labs
360-378-2165
schwinge-at-fhl.washington.edu






From: Jay Jerome :      jjerome-at-bgsm.edu
Date: Tue, 18 Mar 1997 16:23:06 -0500 (EST)
Subject: Re: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Peggy-

In our experience the following are the most important issues for good
antibody staining-

quality of primary ab } fixative } buffer } dilution } blocking agents.

Most of our background problems can be corrected by getting a better
antibody. Some out there work great on gels but once in among all the
cellular proteins they start binding to everything.

The proper fixation (which of course varies for every different antibody
antigen mix we try) is the next most critical.

We have also found differing the buffers can increase the signal to noise
ratio. Although we normally use Gey's salts (it seems to works well with
a whole range of antigens) sometimes cacodylate is superior.

The bottom line, however, is that with any new antigen or antibody it
takes us a lot of trial and error to find the correct match.

Although we have not yet discovered a magic potion, I hope this limited
answer is helpful-

Jay Jerome
**************************************************************
* aka: W. Gray Jerome *
* Dept. of Pathology *
* Bowman Gray School of Medicine of Wake Forest University *
* Medical Center Blvd *
* Winston-Salem, NC 27157-1092 *
* 910-716-4972 *
* jjerome-at-bgsm.edu *
**************************************************************

On Tue, 18 Mar 1997, Peggy Brannigan wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello immunogold experts!
}
} Hope I'm not duplicating an earlier topic but couldn't find anything in
} postings kept over the last few months so, here goes...
}
} What factor (s) in an immungold labeling protocol has (have) the greatest
} impact on background, e.g.. antibody dilution, buffer, etc.? I'm
} looking at the immunogold protocol itself so I can use valuable tissue
} already embedded.
}
} (I'm labeling virus-infected plant leaf tissue with monoclonal antibodies
} to viral proteins - samples are embedded in LR White and fixed in 3%
} paraformaldehyde/0.2% glutaraldehyde. We block before primary antibody
} incubation with a TRIS-HCl buffer containing bovine serum albumin and goat
} normal serum. Our secondary antibody is a commercially prepared goat
} anti-mouse gold conjugate. Usually we have excessive background on
} chloroplasts, mitochondria and cell walls; if we get rid of background by
} increasing dilutions of antibodies, we lose specific reactions as well).
}
}
} One interesting point....we don't have this problem with polyclonal
} antibodies, only monoclonals.
}
} Thanks for any tips -- this is really frustrating!
}
}
} Peggy
}
} Peggy Brannigan
} Electron Microscopy
} Floral and Nursery Plants Research Unit
} National Arboretum
}
} Bldg. 010A R.238
} 10300 Baltimore Avenue
} Beltsville, MD USA20705
}
} Phone: (301) 504-6097
} Fax : (301) 504-5096
} Email: brannign-at-asrr.arsusda. gov
}
}
}
}




From: Jbarjr-at-aol.com
Date: Tue, 18 Mar 1997 17:22:20 -0500 (EST)
Subject: TN5500 with Dec1123+

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello user group,

Im looking for a functioning TN5500 with a Dec PC preferable an 1123 or
better
without the Detector for a project I'm working on. Has any of you recently
upgraded
thier probes and is looking to sell or donate it.
I will gladly pay all associated expenses. Please contact me directly at the
following:
Regards,

Joe Barney
Micro-Analytical Service Center Inc.
Phone: 717-299-0599
Fax: 717-299-2022
E-mail: Jbarjr-at-aol.com




From: tnjed-at-uic.edu (Jack Gibbons)
Date: Tue, 18 Mar 1997 16:46:56 -0600
Subject: yeast morphology

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am doing a tem project with yeast and will have to identify membranes and
organells. Can anyone suggest a good book on yeast morphology?

Sorry about the attachment on the earlier message, it was an error.

Thanks in advance.

Jack Gibbons
tnjed-at-uic.edu






From: RCHIOVETTI-at-aol.com
Date: Tue, 18 Mar 1997 18:04:55 -0500 (EST)
Subject: Re: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Peggy,

One thing which I've found to be helpful for getting rid of non-specific
background labeling is 2-3 washes with a high-salt buffer immediately
following the primary Ab. I've done quite a bit of work in the past with
biotinylated primaries followed by streptavidin-gold, and have seen cases of
high background on occasion; it somehow seems to be primary
Ab-dependent...very bizarre.

Anyway, I'm wondering if something similar might be happening in your system.
If you want to give it a try, I would suggest the following:

1. Make a buffer (TRIS HCl should be fine) containing 5 normal saline.
Regular normal saline is 0.9% by weight, or about 150 mM, so add enough NaCl
to make it 4.5% by weight (about 750 mM). I know this sounds like a whopping
dose of salt, and I can't give a complete rationale for why it works, but it
*does* decrease the background!

2. Do 2-3 washes w/ the high salt following the primary, then a couple of
washes w/ the normal TRIS buffer to get the ionic concentration back where it
should be for the gold conjugate.

3. One other thing: Do all of your diluents contain normal goat serum (even
the diluent for the gold)? If not, you might want to try this as well.

Best of luck!

Bob Chiovetti




From: Daryl Webb :      dwebb-at-waite.adelaide.edu.au
Date: Wed, 19 Mar 1997 10:55:43 +1030 (CDT)
Subject: Re: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Peggy Brannigan writes:
}
} (I'm labeling virus-infected plant leaf tissue with monoclonal antibodies
} We block before primary antibody
} incubation with a TRIS-HCl buffer containing bovine serum albumin and goat
} normal serum.

Peggy, we had similar problems using some home grown antisera. Our
solution was to block in 2%BSA, 5% Normal X serum and 1% Skim milk powder
in TBS for 1 hr. Didnt completely solve the cell wall binding but was a
*huge* improvment, didnt affect the specific binding at all.

Worth a try at least
--

Daryl Webb (dwebb-at-waite.adelaide.edu.au)
Dept. of Plant Science, Waite Institute
University of Adelaide, Glen Osmond S.A. 5064
Australia. Voice:61_8 8303 7426 Fax:61_8 8303 7102







From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Tue, 18 Mar 1997 21:55:19 -1000 (HST)
Subject: Summary of "What Res Scanner?"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Many thanks to everybody who replied to my query about flatbed scanners.
You are all a font of all kinds of information! To summarize:

1) (Obviously) get the highest res scanner with the most pixel depth you
can afford - it's hard to overdo.

2) Printers that are 300-600 dpi with dither and print grey levels at
100-200 dpi; for continuous tome they need 3x3, 4x4, or 9x9
points (depending on the type of printer) for each pixel. So the higher
the resolution of the image scanned in the better the print (also
obvious).

3) Opinions vary - some say that 24 bit color depth gives 256 grey levels
which is "more than enough", while others say that 30 or 36 bit depth
gives more contrast enhancement, even when your software only handles 24
bit color. The best bet is to try out 24 and 30 bit scanners and see
which does the job for you.

4) The big problem lies in enlarging images. A 4x5 SEM negative scanned
in at 300 dpi can really only become a 4x5 print. A TEM negative scanned
in at 600 dpi can be enlarged {3x, whereas a TEM negative can be enlarged
10x photographically. This brings up the discussion at MSA '96 - learn to
take your micrographs at the magnification at which you will want to use
and reproduce them rather than enlarging and therefore gaining only "empty
magnification".

5) Remember to check for the true resolution, not the interpolated res.
of a scanner.

6) Be careful of getting ghosts from glossy Polaroids and Newton rings
from TEM negatives on some scanners.

7) None of the flatbed scanners seem good enough for 35mm negatives - buy
a film scanner instead.

8) Remember that higher res images mean larger files. WIll you be able
to store them? You will soon be needing ZIPs, JAZs, and CD-Rs. (Our SEM
students are alloted 10 MB storage space on their department accounts.
The full-size, full-res, colorized Mexican Ant on my web site is over 8
MB by itself!)

9) See "Tips & Tricks" at http://www.biotech.ufl.edu/~emcl/ for a
previous discussion on this topic.

As you can see from my "cheap frame grabber" and "cheap scanner" posts, we
are trying to operate within an embarrassingly small budget this year.
ALthough I am still fairly happy to prepare everything photographically
for myself, when I talk about image presentation and look at the bright
shining faces of my current SEM students, I know that I am not going to be
teaching them darkroom techniques. They start with Polaroid 55 pos/neg
film. One needs to make slides right away; another wants to know how to
make camera ready illustrations for a journal by next week. Most of these
students ultimately have access to some decent hardware and software, and
I need to point them in the right direction. I need about $10K to bring
my lab up to speed. Too bad I can't invite you all to a huli-huli chicken
sale!

What brought this up? Microtek 300 dpi scanners are down to $195, which I
could buy out of my own pocket. It was tempting.

Aloha,
Tina

http://www.pbrc.hawaii.edu/bemf/microangela
****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************





From: Andreas Loewe :      loewe-at-uni-bonn.de
Date: Wed, 19 Mar 1997 09:56:34 +0200
Subject: Q: Coating of Ceramic Samples for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear folks,

we recently got a new SEM.

What we need now is some advice for coating ceramic samples
When to use Carbon, Gold, thickness etc.

Is there any FAQ about this available?

Any help highy appreciated.

______________________________________________________________
Andreas Loewe Tel: +49-228-734-180
University of Bonn Fax: +49-228-734-205
Insitute for Inorganic Chemistry email: loewe-at-uni-bonn.de
Inorganic Material Research
Roemerstr. 164
53117 Bonn
Germany http://www.elmi.uni-bonn.de/
______________________________________________________________






From: Peggy Brannigan
Date: 18 March 1997 18:39
Subject: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Peggy

your problem might be quite fundamental. It's several years since I've been
seriously involved with immunogold labelling but I recall that there is a
basic problem with monoclonal antibodies.

If the antigenic site, which the monoclonal antibody is specific to, is
particularly sensitive to fixation then it will not label, hence the lack of
specific labelling. You mentioned that a polyclonal works and this could
simply be because it can 'find' a less sensitive part of the antigen. Have
you had a chance to try cryo or less fixation?

I apologise if my comments are out of date or irrelevant.

Malcolm Haswell
University of Sunderland
UK
----------

Hello immunogold experts!

Hope I'm not duplicating an earlier topic but couldn't find anything in
postings kept over the last few months so, here goes...

What factor (s) in an immungold labeling protocol has (have) the greatest
impact on background, e.g.. antibody dilution, buffer, etc.? I'm looking
at the immunogold protocol itself so I can use valuable tissue
already embedded.

(I'm labeling virus-infected plant leaf tissue with monoclonal antibodies to
viral proteins - samples are embedded in LR White and fixed in 3%
paraformaldehyde/0.2% glutaraldehyde. We block before primary antibody
incubation with a TRIS-HCl buffer containing bovine serum albumin and goat
normal serum. Our secondary antibody is a commercially prepared goat
anti-mouse gold conjugate. Usually we have excessive background on
chloroplasts, mitochondria and cell walls; if we get rid of background by
increasing dilutions of antibodies, we lose specific reactions as well).

One interesting point....we don't have this problem with polyclonal
antibodies, only monoclonals.

Thanks for any tips -- this is really frustrating!

Peggy

Peggy Brannigan
Electron Microscopy
Floral and Nursery Plants Research Unit
National Arboretum

Bldg. 010A R.238
10300 Baltimore Avenue
Beltsville, MD USA20705

Phone: (301) 504-6097
Fax : (301) 504-5096
Email: brannign-at-asrr.arsusda. gov





From: Woody.N.White-at-mcdermott.com
Date: 3/19/97 9:56 AM
Subject: Q: Coating of Ceramic Samples for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The general rule:

For x-ray analysis work use carbon (evaporate). It will (typically)
attenuate less than Au and will not generate interfering x-ray lines.
Because of the low Z coating and usual low z of ceramics, imaging of
carbon coated ceramics should be done with as low a beam voltages as
practical.

For the best imaging, use Au (sputter). It has better stopping power
(= better resolution) and will provide a higher SE yield than carbon.
Lower beam voltages (ie: 5-10 kv) still helpful.

Thickness: Always use the thinest coating which will prevent
charging. This can sometimes be a problem on low density, friable
ceramic fractures or fiberous material as a continuously conductive
films can be difficult to achieve. One way to mimimize this is to
paint as much of the specimen as possible (those areas not viewed)
with carbon paint. Wicking can be a problem - viscosity and
volitility of the paint will make quite a difference.

Alternatives (which I do not have available with my system)....

Very low kV imaging....in the area of approx. 500 to 1500 beam volts,
some materials will be at an equilibrium state where energy in =
energy out and thus dont't charge.

E-SEM - High chamber pressure discharges specimen and coating is not

needed. It has been my observation that this will work well for some

applications, but is surely no "cure-all". High mag, high resolution,
low
kV exams usually suffer.

Woody White

______________________________ Reply Separator
_________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Dear folks,

we recently got a new SEM.

What we need now is some advice for coating ceramic samples
When to use Carbon, Gold, thickness etc.

Is there any FAQ about this available?

Any help highy appreciated.

______________________________________________________________
Andreas Loewe Tel: +49-228-734-180
University of Bonn Fax: +49-228-734-205
Insitute for Inorganic Chemistry email: loewe-at-uni-bonn.de
Inorganic Material Research
Roemerstr. 164
53117 Bonn
Germany http://www.elmi.uni-bonn.de/
______________________________________________________________




From: Marti, Jordi :      MartiJ-at-MTOMP201.Research.Allied.com
Date: Wed, 19 Mar 97 08:52:00 EST
Subject: RE: Q: Coating of Ceramic Samples for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html








Andreas wrote:

What we need now is some advice for coating ceramic samples
When to use Carbon, Gold, thickness etc.


Andreas, it will depend ,to some extent, on what you are after. If you
intend to do EDS for example, then you might want to deposit C since Au
could interfere with your results. Also keep in mind that Au will deposit
as relatively large islands and this will limit your resolution. You will be
better off using Au-Pd for example as opposed to just Au. How thick a film
you need will also depend , among other things, on the voltage that you are
using in the SEM and on how flat your sample is (i.e. are you looking at a
fracture surface or a polished surface ?). In most cases, at 10 or 20KeV ,
10-30 nm of Au-Pd should be enough provided the sample is grounded properly
(i.e. use a conductive paint, such as gold or carbon, around your sample).
If you operate at lower voltages , a thinner coating might be O.K. You
might want to read some standard texts on SEM and experiment a bit with
your samples.

Jordi Marti
______________________________________________________________
Andreas Loewe Tel: +49-228-734-180
University of Bonn Fax: +49-228-734-205
Insitute for Inorganic Chemistry email: loewe-at-uni-bonn.de
Inorganic Material Research
Roemerstr. 164
53117 Bonn
Germany http://www.elmi.uni-bonn.de/
______________________________________________________________






From: greg :      greg-at-umic.sunysb.edu
Date: Wed, 19 Mar 1997 09:24:59 +0000
Subject: Re: Q: Coating of Ceramic Samples for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Dear folks,
}
} we recently got a new SEM.
}
} What we need now is some advice for coating ceramic samples
} When to use Carbon, Gold, thickness etc.
}
} Is there any FAQ about this available?
}
} Any help highy appreciated.
}
} ______________________________________________________________
} Andreas Loewe Tel: +49-228-734-180
} University of Bonn Fax: +49-228-734-205
} Insitute for Inorganic Chemistry email: loewe-at-uni-bonn.de
} Inorganic Material Research
} Roemerstr. 164
} 53117 Bonn
} Germany http://www.elmi.uni-bonn.de/
} ______________________________________________________________
}
} Andreas,
First make sure that the sample is well grounded. Then it
will be trial and error as to how thick to coat the sample.
Gold usually gives the best results.--Good Luck

}
Gregory Rudomen
Greg-at-UMIC.SUNYSB.EDU
University Microscopy Imaging Center
S.U.N.Y. Stony Brook




From: Aurion-at-Company.DiVa.nl (Jan L.M. Leunissen)
Date: Wed, 19 Mar 1997 17:04:50 +0100
Subject: Re: Au label - high background

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Peggy,

The undesired signals are present only when you include your monclonal
antibody, if I understand correctly.
I don't think that it is just the fact that you have a monoclonal, after
all, antiserum is more or less merely a collection of monoclonals.

If this undesired signal is caused by an unspecific reaction at the primary
antibody level, as often found e.g. with IgM type antibodies (the bigger
the stickier), you would expect to be able to get rid of it in the firat
place by improving the block step. This takes care of background caused by
hydrophobic interactions.
Secondly, the incubation buffer should have additives (like BSA, cold water
fish gelatin and there is a lot more!) which compete with primary
antibodies for unspecific binding. In that case the protein additive should
prevent or at least reduce background.
Sometimes Tween does help, but since it is a detergent I would only use it
on embedded material, not on cryosections. If you have to use detergent
the best results are obtained at concentrations slightly higher than the
critical micelle concentration.

On the other hand, from your description I get the impression that with
increasing the antibody dilution both your specific and undesired signal
decrease, which may indicate that "both types" of reaction have similar
affinities. The question now is: is this undesired signal background or a
more or less specific signal? Your specific reaction should at least have a
lower Kd than the background reactions. If that is not the case, only a
different antibody with higher affinity will help.

BTW: The topic of specificity and background will be extensively dealt with
at the immunocytochemistry workshop in Iowa City, announced earlier on this
Listserver.

Good luck!

=============================
Jan Leunissen, Ph.D.
AURION ImmunoGold Reagents & Accessories
Costerweg 5, 6702 AA Wageningen
The Netherlands

phone (31)-317-497676
fax (31)-317-415955






From: Bruce Cutler :      BCutler-at-eureka.chem.ukans.edu
Date: Wed, 19 Mar 1997 10:55:43 -0500 (CDT)
Subject: tough embedding

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

To all:

In 1992 Lindley (Microsc. Res. & Techn. 21:355) published a technique
for using a primer (Z-6040, a silane) to allow for good adherence of
biological materials with outer surfaces that do not normally stick well to
embedding plastics. The plastic used was LRWhite. I am interested in
using this primer with Epon 812 substitutes. Has anyone used this
primer with the epoxies and is there a protocol.

In a similar vein, does anyone have a good protocol for embedding cell cultures on
transwell membranes with small pores, such that when sectioning the
cells + membrane the membrane does not rip away from the cells tearing
off the basal layer of the bottom cells. Different membrane polymer
types differ have different effects, as well as different pore sizes.
If a very thick collagen layer is laid on top of the membrane this
helps, but for experimental work this is often undesirable.

Bruce Cutler, Microscopy Laboratory, Univ. Kansas, Lawrence,
KS 66045-2106, (913) 864 - 4140






From: Ricky L Vaughn :      RLVAUGHN-at-MAIL.UNMC.EDU
Date: Wed, 19 Mar 1997 11:39:28 -0600
Subject: re slide film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I just had to do a job for an investigator that wanted their prints made into
slides. I have usually let them them take care of that and found out that
our Biomed's graphic people use color slide film only which heard gives a
blue cast to the EM prints. I decided to try Kodak's RPC # 175-3151 film on
our copy stand (the people I got it from use it for taking slides of X-rays).
It turned out Great! The draw back.... 10 and 15 second exposure
brackets, at f4 f-stop, with four 500 watt photofloods. I developed it in
our Xomat x-ray processor but was told you can use D-19 also. I'm going
to reshoot the prints with Fugichrome 100 and that Direct MP 5360
someone else mentioned to compare. Has anyone else done this
comparison? My next trick is to find out how to get the Poloroid Digital
Pallete we have to use these long exposure films, if possible.

Rick Vaughn
Electron Microscopy Research Facility




From: DON_STEELE-at-CCKRDC.CA.ALCAN.CA
Date: Wed, 19 Mar 1997 10:40 -0500 (EST)
Subject: Flatbed scanners and "moire"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to all who responded to my cry for help recently re: what I
mistakenly was calling moire patterns on images scanned using my Agfa
Arcus II scanners.

Turns out, as was recently mentioned in a posting by Tina Calvaho
(sorry if I got that wrong Tina), that the the effect is attributed to
"Newton Rings" by those in the biz. The problem shows up when
scanning negs which don't have very much detail.

The primary culprit in my case was a misalignment of the transparency
module (ie the lid was on crooked). The effect was most prominent on
the scanner with the greatest misalignment. This is not something
which you can adjust unfortunately (holes in the chasis which accept
the mounting brackets were slightly off), but Agfa was quick to send
me another unit with it's head on straight.

None of my units are "perfect", but I'm told that something called
"Newton Ring Spray" or matte spray solves any remaining problem. I'm
hesitant to spray something on the glass, but if it's removable, I'll
likely give it a try.

Cheers,

****************************************
Don Steele Steele-at-KRDC.INT.Alcan.Ca
Alcan International
Kingston Research and Development Centre
(613) 541 - 2145
****************************************






From: colijn.1-at-osu.edu (Henk Colijn)
Date: Wed, 19 Mar 1997 15:18:31 -0500
Subject: Re: slide film & TEM negative scanners

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

{snip...}

} comparison? My next trick is to find out how to get the Poloroid Digital
} Pallete we have to use these long exposure films, if possible.


Rick brings up a point that should be considered when evaluating a scanner.
Not only should you consider the pixel depth (8, 10 or 12 bits), but you
should also consider the optical density range of the scanner. I don't
think the recent thread on scanners touched on this.

The digitization depth basically indicates the *output* range of the
scanner. A 10 bit scanner will output ~1000 gray levels. The optical
density range, though, indicates the *input* range of your scanner.
Optical density (OD) is the log(10) of the fraction of transmitted light.
Since most scanners start at roughly an OD of 0, a scanner with an OD range
of 2 will be able to "see" down to about 1% light transmission. An OD
range of 3 provides you with a range of 1000 in the light transmission.
With the same number of bits, each step will be 10 times coarser. You
will, however, be able to pull information from those dark regions of the
negative. Consider what kind of negatives you will be scanning. If they
are relatively flat, an OD range of 2 may be sufficient. If you have high
contrast negatives, check whether you need the extra input range.

The Umax scanner I've used has an optical density range of ~2.0. Agfa
indicates that their Arcus II (US$1500-2000) has a density range of 3.0
(0.2 - 3.2). Their DuoScan ($4000-4500) has an OD range of 3.3 (0.2 -
3.5?). Polaroid's Sprintscan 45 apparently has an OD range of 3.4 with a
$10k list price (ouch!).

So... think about your OD needs as well as pixel density, digitization
bits, and price!

Cheers,
Henk




Henk Colijn colijn.1-at-osu.edu OSU Campus Electron Optics Facility
-------------------------------------------------------------------
Prosperity is the blessing of the Old Testament; Adversity is the blessing
of the New. Francis Bacon, "Of Adversity."






From: pat hales :      hales-at-medcor.mcgill.ca
Date: Wed, 19 Mar 1997 16:05:42 -0800
Subject: histochemical stains for plastic sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Other than toluidine blue, as you mentioned, we also use a variation on the
hematoxylin theme for our epon sections. We use Heidenhain hematoxylin-iron
which comes in two solutions (I have the recipes if you want to make your
own) - an iron alum solution and a hematoxylin solution. Slides are
preheated and stained on a hot plate that is maintained between 80 and 90
degrees C. Coat the sections with the iron alum solution for 2-10 min
depending on the tissue type, thickness, etc. Rinse VERY WELL with distilled
water. Repeat the process on the hot plate with the hematoxylin staining
solution using the same time interval as the iron alum. Because the
"staining" only appears at this step, you may need to test a few slides to
determine ideal timing. After rinsing well again with distilled water, flood
the slides on the hot plate with TAP water to differentiate and let sit for
3 min. Rinse briefly in distilled water and dry. Prestaining like this
allows the slides to then go through autoradiography without interfering
with the emulsion and has worked for us irregardless of the fixative used.

Pat Hales
McGill University
Dept. of Anatomy & Cell Biology
hales-at-hippo.medcor.mcgill.ca






From: R. Edelmann :      edelmare-at-CASMAIL.MUOHIO.EDU
Date: Wed, 19 Mar 1997 16:47:09 -500
Subject: More info regarding Denkil SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for some one who as any information on something called
antistatic "Denkil SEM" (Hodogaya Chemical Co. Ltd.). It was cited
in a short article in J. Electron Microsc, Vol 28, No. 4 P 312-313.

Hoping to spark some ideas I have included the article, sic. (I apologize
for any inconvience for members who are charged by file size basis,
but its quite short:). Please, excuse any typos as I am not a
professional typist, but I haven't altered any wording. Sorry I
haven'y included the figures (but they do look impressive enough to
bother trying to follow up)

----+----+----+----+----+----+----+----

J. Electron Microscopy, 1979, Vol. 28, No.4, 312-313.


Scanning Electron Microscopy of Lily Pollen Grains treated with
Antistatic Solution

Masaru Katoh

(Nissei Sangyo Co. Ltd.)


Pollen grains are generally observed by SEM after drying in air and
coating with metal. in some special pollens or in studies with
higher resolution and magnification, however, routine procedures for
preparation of general biological materials may be necessary.

In this letter, effects of acohol-soluble antistatic agent (1) upon
electroconductivity of the specimen of lily pollen grains are
introduced. in this method, specimens do not need troublesome metal
coating and preserve much more natural shapes.

Figure 1a shows pollen grains of lily simply dispersed on a small
piece of adhesive tape without following procedure of metal coating.
in this micrograph, a trouble of charging up of the specimen is
recognized. figure 1b is a micrograph from the same area as shown in
Fig. 1a, but was taken after the treatment that the specimen drawn
out from the microscope was immersed in several drops of 3% alcohol
solution of the antistatic "Denkil-SEM" (Hodogaya Chemical Co. Ltd.,
Japan), and throughly dried in air. by this treatment, charging up
of the specimen was completely eliminated. Another important effect
of this method is that the pollen grains are expanded and the net
pattern of the surface structure is restored. This may be a big
advantage of this method keeping the pollen grains closer to the
natural shape. This effect is probably due to a strong affinity of
the antistatic agent with water restoring the orginal shapes of the
pollen grains. Effective water may be derived from the dissolving
component in alcohol and humid atmosphere. the absorbed water is
probably retained firmly in the pollen grains even after exposuring
to electron beam in a high vacuum.

Such an electroconductive treatment using water-retainable
antistatic agent is considered very useful not only for prevention of
charging up of specimens, but also, to some extent, for keeping some
specimens moist even in a high vacuum.

Further examples of application of this antistatic agent will be
reported in the near future together with many other interesting
results.


Reference:

(1) Katoh, M. : J. Electron Microsc., 28, 51 (1979)


-----+------+------+-------+-------+-------


That's it thats the entire paper. Does anyone out there have any
ideas to point us in the right direction?

Richard E. Edelmann, Ph.D.
Electron Microscopy Facility Supervisor
352 Pearson Hall
Miami University, Oxford, OH 45056
Ph: 513-529-5712 Fax: 513-529-4243
E-mail: edelmare-at-muohio.edu




From: Doug Keene :      DRK-at-shcc.org
Date: Thu, 20 Mar 1997 23:30:52 -0600 (cst)
Subject: slide films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear slide makers:

I've noticed a great deal of discussion regarding films for
slides, and given how important this topic is to
microscopists, I thought I'd add my experience to the list.
My original film was also the MP 5360. The film works
well, is of high resolution, and exposure times of one
second or less using a MP-4 copy stand are common. You
can process it in your own lab. The drawback, in my
opinion, is a yellowish-brown hugh in the background. I am
told this can be cured with fresh bulbs in the MP4 copy
stand, and that they should be replaced "routinely", but I
really do not want to be bothered with that responsibility
and expense. I have also used the polaroid instant slide
film, but I find the resolution unacceptable, and the
slides I have made and archived do not seem to hold up to
the test of time. Our current solution? We use Kodak
Ektachrome 64T ("T" for tungsten lighting), EPY 135-36
film. The resolution is top shelf, and exposure times are
on the order of 1 second or less. Standard E-6 processing
at the local film processing stand does a nice job, so you
don't even have to get your hands wet. On a 35mm camera
set up on the MP4 copy stand, we use a film speed of "50",
and f-stop of 2.8, and automatic exposure. I do not know
the specifics about the lens which we use, but the distance
to the camera is from about 8 inches to 2.5 feet, depending
on the enlargement desired. With this film you will also
experience a brownish background. Using a SLIGHTLY blue
filter over the camera lens (available at the camera shop),
a pleasing grey background can be gained. As a darker blue
lens is used, the slides will become increasingly blue, but
I for one prefer blue over brown.

Hope this helps,

Doug Keene
Shriners Hospital for Children
Portland, Oregon
----------------------
Doug Keene
DRK-at-shcc.org






From: bjg-at-cyllene.uwa.edu.au (Brendon J. Griffin)
Date: Thu, 20 Mar 1997 11:37:58 +0800
Subject: ?Lines on laserwriters - Tektronics Phaser 550

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear all,

I recently purchased the Tektronics Phaser 550 colour laserwriter with
extended features (1200 dpi) at great cost - ~$12K (OZ$). It arrived last
week to great joy but test prints showed a regular fine horizontal line
repeated at a spacing of 26.1mm when printing at premium grade. The
spacing is 35mm at all other grades. The line is ~0.1-2 mm in width. It
occurs in all images, ie colour, B/W and on test colour stripe images, ie
it is independant of image type or source.

Complaint led to a rapid visit by the service engineer who confirmed it as
a problem and that a second complaint had been lodged from a similar new
installation several days earlier.

The Company response is now - that's laserprinters! IE the Tektronics
Phaser 550 will always produce this regular and quality destroying line. I
agree it is not strong but once noticed it catches the eye.

My query is:

has anyone else observed this regular lining?

Has anyone got a Phaser 550 which does not produce the lines?

I feel that this is totally unacceptable, needless to say these lines were
not evident in sample prints, but the overall quality is good and the media
print pricing is also good.

Note I am aware of dye-sub etc, I was after a unit with a lower media unit cost.

Thank you


Brendon J. Griffin
Centre for Microscopy and Microanalysis
The University of Western Australia
Nedlands, WA, AUSTRALIA 6907
ph 61-9-380-2739 fax 61-9-380-1087

*Please note change to email address:
bjg-at-cyllene.uwa.edu.au






From: Mary Mager :      mager-at-unixg.ubc.ca
Date: Wed, 19 Mar 1997 22:38:33 -0800
Subject: Re: Q: Coating of Ceramic Samples for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Andreas,
The simple answer is:
1.) Use carbon evapoative coating when you want to do an EDX analysis of the
ceramic. The thickness of carbon when a polished brass disk beside your
sample just turns blue is 25 nm and is adequate for most samples.
2.) Use sputtered gold or sputtered 60%gold - 40%palladium alloy when
imaging is your goal. Images are better when the sample is gold sputtered. A
10 nm coat is good. Your sputter coater should have a time vs. coating
thickness chart.
You wrote:
} we recently got a new SEM.
}
} What we need now is some advice for coating ceramic samples
} When to use Carbon, Gold, thickness etc.
}
} Is there any FAQ about this available?
}
} Any help highy appreciated.
Good luck with your new SEM.
Regards,
Mary
Mary Mager
Electron Microscopist
Metals and Materials Eng., UBC
6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel:604-822-5648, fax:604-822-3619
e-mail: mager-at-unixg.ubc.ca





From: Braun :      braun-at-argos.ipfdd.de
Date: Thu, 20 Mar 1997 11:02:48 -0500
Subject: JobVacancy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Job vcancy for PhD student / Research Cooperation within area of
biomimetic chemistry

We are looking for a PhD student to work on a project focusing on
" Chemical reactions at microstructured surfaces ".
The project is in cooperation with an industrial partner and
the Central German Research Agency
( Leitz Electron Optics , LEO ) and focuses on the
- chemical surface modification by microprinting techniques as
published by Whitesites et. al.
- study of surface reactions (for example formation of inorganic
phases at patterned surfaces)
within patterned surfaces in order to create spatially defined
product areas.
The work follows the ideas of biomimetic reactions at organic
(polymeric) surface layers.

The structural part of the work focuses on the
structural characterisation of surfaces and the reaction products
formed at the surface. This should be done by microscopic methods
inlcuding high resolution field-emisson scanning electron microscopy
as well as Energy spectroscopic imaging and EELS analysis.

Experience in the surface modification by micropatterning is
already available.

The project lasts till 31.12.1999

A cooperation with groups engaged in surface reactions, biomimetical
reactions could also be possible. (Exchange of PhD students within
the frame of this project).

For further information contact :

Dr. H.-G. Braun
Institute of Polymer Research Dresden
Microstructure Groups
D-01069 Dresden
Hohe Str. 6
Germany

Mail : Braun-at-argos.ipfdd.de




From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Thu, 20 Mar 1997 08:00:20 -0500
Subject: Re: ?Lines on laserwriters - Tektronics Phaser 550

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have experienced the same with a Lexmark Optra R at the higher dpi
settings. Company says that it will always be a problem because the
manufacturing process allows a rather high tolerance in the production of
the gearing leading to chatter. They have given us two other machines which
show the same problems. Needless to say we are quite dissatisfied with the
product.




At 11:37 AM 3/20/97 +0800, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {
GO GATORS
Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 352-392-1295
ICBR EM Core Lab fax 352-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/
The home of " Tips & Tricks "









From: Sandra F. Zane :      sfzane-at-unccvm.uncc.edu
Date: Thu, 20 Mar 1997 10:43:12 -0600
Subject: Direct Positive Microfilm 2468

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello Everyone,
There has been a tremendous response to my plea for information
about the DPM 2468. At the request of a couple of you, I am posting my
findings to the list. Here is basically what I have learned from you folks,
Kodak, and a couple of dealers with whom I have spoken.

1. Kodak is still making the DPM 2468. They won't produce the
film, however, until they have an order for it.

2. Kodak will not sell this film in quantities less than 50 100ft.
rolls. This makes dealers reluctant to stock it.

3. Neelima Shah informed me that Mid City Camera in Philadelphia
carried this film and I was able to order 2 rolls from them.
However, they are not sure that they will continue to stock it and
are going to call me later this week to let me know what they
decide. The 2 rolls they sold to me were the last they had in
stock.

Now for the good news: Several of you have informed me of another
film which seems to be very similar to this one in ease of use, exposure
times and development. It is called Direct MP Film 5360 and is available
from Ted Pella. You can find it on page 186 of their catalog. Those of you
who have used this film seem to be pleased with the product.
I would like to thank all of you who responded to my post on this
listserver and for the information you shared with me. I attempted to
answer each of you individually; but because the response became so great, I
will have to thank you as a group. Once again you were there and willing to
help.
I am not affiliated with any of the organizations who make or sell
any of the products discussed here.
Again, thanks to all who shared. Sandra

Sandra F. Zane, EM Tech. sfzane-at-email.uncc.edu
Dept. of Biology, UNCC Ph.(704)547-4051
9201 University City Blvd. Fax (704)547-3128
Charlotte, NC 28223





From: pat hales :      hales-at-medcor.mcgill.ca
Date: Thu, 20 Mar 1997 14:50:32 -0800
Subject: UV Embedding in Flat Molds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am currently trying to switch from Epon embedding media to one which
polymerizes by UV at lower temperatures. In order to orient my tissue
properly I would like to continue using flat embedding molds. My problem is
in making these molds "airtight" in order to prevent shrinkage/evaporation
while it polymerizes. Has anyone had this problem or found a solution to it?

Pat Hales
McGill University
Dept. of Anatomy & Cell Biology
hales-at-hippo.medcor.mcgill.ca





From: MARK DARUS (216) 266-2895 GENERAL ELECTRIC CO. :      darus-at-cle.dnet.ge.com
Date: Thu, 20 Mar 97 15:19:51 EST
Subject: Triple Carbonate Solution

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have a Voyager EDX system. I want to perform a standardless semi-quant
analysis on a solid solution. The solid solution is a combination of Barium,
Strontium and Calcium carbonates, (BaCO3, SrCO3 & CaCO3). Can I have the
Voyager look at the Ba, Sr & Ca and tell it that these exist as carbonates
and have it calculate the weight percentage of each from me entering in the
stoichiometry of the solution?
Are there other suggestions for me to go about measuring the weight
percentages of Barium carbonate, strontium carbonate and calcium carbonate in
a solid solution?




From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Thu, 20 Mar 1997 16:24:19 -0500 (EST)
Subject: Re: UV Embedding in Flat Molds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Date: Thu, 20 Mar 1997 16:32:07
} To: pat hales {hales-at-medcor.mcgill.ca}
} From: Scott Whittaker {sdw-at-biotech.ufl.edu}
} Subject: Re: UV Embedding in Flat Molds
}
}
}
} A procedure we use around here successfully is a piece of cleared film ( or
aclar) punched with a hole puncher and placed on a pre-polymerized resin
surface in a centrifuge tube. Just drop in the film, add tissue, add the
resin, and polymerize as you would normally.
} After curing, remove the tube and break the sample at the disk level. The
resin doesn't stick to the film (or aclar) and you are left with the tissue
at the surface of the block and ready to section in any orientation. Good luck
}
}
}
}
}
}
} At 02:50 PM 3/20/97 -0800, you wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {
GO GATORS
Scott D. Whittaker 218 Carr Hall
Research Assistant Gainesville, FL 32610
University Of Florida ph 352-392-1295
ICBR EM Core Lab fax 352-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/
The home of " Tips & Tricks "









From: Robert Mixon :      mixonr-at-ohsu.edu
Date: Thu, 20 Mar 1997 13:52:47 -0800
Subject: PNEMS Spring Meeting, Portland OR

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The Pacific Northwest Electron Microscopy Society wishes to extend a warm
welcome to all who wish to attend the Spring '97 Meeting at the V.A.Medical
Center in Portland, Oregon Friday May 2nd, 1997. There are still a few spots
for the up and coming young investigators/ graduate students to present papers
on their current work to an enthusiastic Northwest audience of colleagues.
Please send an e-mail to Charlie Meshul, Program chairman at meshulc-at-ohsu.edu
with your topic if you wish to added on. Alternatively send e-mail to Bob
Kayton kayton-at-ohsu.edu or Bob Mixon mixonr-at-osu.edu

THANKS




From: Doug Keene :      DRK-at-shcc.org
Date: Fri, 21 Mar 1997 21:33:44 -0600 (cst)
Subject: conjugation of primary ABs to gold particulates

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Has anyone hired a commercial enterprise to conjugate
their primary antibody to a gold particulate? We have
a couple of mouse monoclonals that we would like to
simultaneously localize in tissue, but do not have
the time or experience to do the conjugations
ourselves.

Many thanks,

Doug Keene
Associate Investigator
Shriners Hospital for Children

----------------------
Doug Keene
DRK-at-shcc.org






From: Tobias Baskin :      baskin-at-biosci.mbp.missouri.edu
Date: Thu, 20 Mar 1997 16:11:16 -0600
Subject: Re: UV Embedding in Flat Molds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Greetings,
For flat embedding under uv light without air, we use flat bottom
capsules. These are made by TAAB in UK and distributed by several USA em
suppliers. If you have really big samples, I beieve the little plastic
widgets that em grids come in will work and these too are available from
suppliers (EMS at least, and no doubt others too).
Hope this helps,

Tobias Baskin

_ ____ ^ __ ____ Tobias I. Baskin
/ \ / / \ / \ \ University ofMissouri
/ | / / \ \ \ BiologicalSciences
/___/ /__ /___ \ \ \__ 109 Tucker Hall
/ / / \ \ \ Columbia, MO 65211 USA
/ / / \ \ \ voice: 573-882-0173
/ /____ / \ \__/ \____ fax: 573-882-0123






From: DUNNTEM-at-aol.com
Date: Thu, 20 Mar 1997 17:09:54 -0500 (EST)
Subject: color 35mm neg scanner

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have hundreds of 35mm color negs which I wish to make proof sheets of.
I do not want to do anything with those images other than print proof sheets.
Can anyone suggest a good scanner for this job.

Thank you.
Ted Dunn




From: lizard-at-okway.okstate.edu (Ginger Baker)
Date: Thu, 20 Mar 1997 13:33:49 -0600
Subject: preparation/coating of beeswax

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello all,

I have a PhD student from Brazil who is studying Catolaccus
grandis. He places a female on top of a beeswax cell which contains
boll weevil larva. The female deposits an egg on the larva and 2
weeks later, it matures and burrows its way out of the beeswax
creating a hole. He would like to study and measure these holes.

However, from past experience I know that beeswax melts when placed
in my Au/Pd sputter coater. Is there any way to coat the beeswax and
examine them with an SEM? Or should he try another route?

Thank you in advance,

Ginger Baker
EM Lab Manager
Dept. Anatomy, Pathology, and Pharmacology
250 Veterinary Medicine
Oklahoma State University
Stillwater, OK 74078
(405) 744-6765
FAX: (405) 744-5275
Email: lizard-at-okway.okstate.edu





From: Doug Keene :      DRK-at-shcc.org
Date: Fri, 21 Mar 1997 23:42:49 -0600 (cst)
Subject: Flat embedding of LR White, Lowicryl

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Pat:

Regarding your question about flat embedding samples in
media which polymerize under U.V.. We have used both LR
White and Lowicryl resins successfully for flat embedding.
We use Beem flat embedding molds, which are quite stiff and
seem to contain less free oxygen than standard molds. Our
trick to exclude atmospheric oxygen is to use the plastic
pipette tip box that "Multi-Flex" tips are packaged in. It
is a three part box which consists of a UV light penetrable
top, which fits over a multi-holed mid section (normally
supporting the tips) which fits over a lower compartment.
I collected a bunch of these from another lab, but I do not
know where to get them. We fill the lower compartment
with dry ice (solid CO2), and drill a couple of holes in the
top (letting CO2 gas escape as the solid sublimes). The
samples in the beem flat embedding molds are placed on the
perforated "shelf" above the CO2 chamber and exposed to a
constantly replenished environment of gaseous CO2, excluding
incoming oxygen. We put the whole thing, including a UV
light, in the -20 portion of our freezer. The
polymerization proceeds fully enough by the time the dry ice
is gone to prevent oxygen inhibition. You might also try
using a "heat sink" type of mold. Ted Pella sells one using
a teflon mold which slides over aluminum ( I have no
commercial interest in Pella).

Good luck,

Doug Keene
Shriners Hospital EM Lab
Portland, Oregon
----------------------
Doug Keene
DRK-at-shcc.org






From: RCHIOVETTI-at-aol.com
Date: Thu, 20 Mar 1997 19:28:40 -0500 (EST)
Subject: Re: UV Embedding in Flat Molds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Pat,

You didn't mention what kind of UV-polymerizing resin you'll be using, but if
it is Lowicryl or a similar methacrylate, here's how to do it:

1. Use polyethylene flat embedding molds. These molds are made from the
same type of material as BEEM (R) capsules. The polyethylene molds are
transparent to the UV light, making the polymerization much more even. Get
back to me if you can't find a vendor for the polyethylene molds, I can't
remember at this moment who sells them -- Ted Pella, I think...but I can look
it up for you.

2. Place the mold on a piece of cardboard that is wrapped in aluminum foil.
The cardboard should be larger than the mold. This makes moving and
transferring the mold much easier and less messy (see below).

3. Overfill all of the cavities in the mold to give a "positive meniscus" at
each position. If some of the cavities don't have specimens in them, go
ahead and fill the cavities anyway.

4. Cut a piece of Parafilm (R) that is slightly larger than the mold. Then,
beginning at one end of the mold, gently lay the Parafilm down on top of the
mold. As you do this, the excess resin will run out of the mold cavities and
get trapped under the Parafilm. Gradually lower the Parafilm down toward the
other end of the mold, allowing the resin to fill the space between the mold
and the Parafilm. This seals the complete top of the mold. The resin won't
dissolve the Parafilm, but it will make it soft.

5. You can use tissues to absorb the excess resin as it runs down the
outside of the mold. (**CAUTION:** Be sure to wear gloves at this step, and
at all steps when working with methacrylate resins -- or any resin, for that
matter!).

6. Transfer the filled molds to your polymerization apparatus. I don't know
what you will be using, but I make polymerization chambers out of cardboard
boxes. Just make two cutouts: one on the top for the UV lamp, and one on
the side for a door. **To make the polymerization even, cover all of the
inner surfaces of the box with aluminum foil.**

7. **NOTE**: You can also use Thermanox (plastic) cover slips instead of
Parafilm in Step #4. Just lay the coverslips on top of the cavities, letting
the excess resin run out along the edges.

8. Let everything equilibrate at low temperature for 15-20 minutes, turn on
the UV lamp, and you're on your way to polymerized resin!

9. **NOTE**: If you see "bubbles" or vortex-like "swirls" in the
polymerized blocks, this is usually a sign that the polymerization was too
rapid. Increase the distance between the lamp and the mold to cut down on
the UV intensity. If this isn't possible, place a piece of frosted glass
between the lamp and the molds.

Hope this helps!

Bob Chiovetti




From: RCHIOVETTI-at-aol.com
Date: Thu, 20 Mar 1997 22:54:21 -0500 (EST)
Subject: Re: conjugation of primary ABs to gold particulates

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Doug,

I've never asked anyone to conjugate a monoclonal Ab directly to colloidal
gold, mainly because of the added cost (a similar concern of yours, I take
it!).

Are you up for a slightly different approach, which you can do with
commercially available reagents? If so, you could do the following:

1. On one side of the grid (maybe a fairly high hexagonal mesh, uncoated
nickel grid?) incubate with one of your mouse monoclonals.

2. Follow with biotinylated goat anti-mouse Ig-whatever (is it an IgG?).

3. Follow with streptavidin-gold (available in many different sizes).

4. Flip the grid over and repeat the labeling with the second monoclonal.
Only this time, use a different size of streptavidin-gold.

5. Double sided labeling works quite nicely for co-localizations, and it
eliminates a lot of cross-reactivity problems you'd have to fight if you
worked on only one surface of the section.

6. One thing to watch out for: If you add normal goat serum and maybe a
little TWEEN or other additives to the reagents, the grids will tend to sink
if they're incubated on drops of the reagents. Then you're labeling *both*
sides with the *same* reagents. This, of course, is to be avoided! You have
to keep the sides separated.

The best way to avoid this problem is to anchor the grid by its edge on a
piece of double-stick tape on a microscope slide. You can then use a
micropipette to deliver *very small* volumes (maybe 2-3 microliters) of each
reagent. To change reagents or to wash, simply add the reagent to one edge
of the grid and wick it off with filter paper
from the opposite edge.

Following the first gold incubation you can remove the grid and wash well in
a gentle stream or w/ repeated drops of buffer and finally water. Allow to
dry, flip the grid over, stick it down to the tape and do the second
run-through.

An added bonus of this procedure is a fairly good signal amplification. If
you conjugated your primary directly to the gold, you'd only have a one-step
reaction with minimal gold particles (probably 1) binding at each site.
Would that be enough to detect above background?

Unless, of course, you *want* the most specific localization possible...In
that case, forget everything I've just written!

Good luck!

Bob Chiovetti




From: RCHIOVETTI-at-aol.com
Date: Thu, 20 Mar 1997 22:54:21 -0500 (EST)
Subject: Re: conjugation of primary ABs to gold particulates

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Doug,

I've never asked anyone to conjugate a monoclonal Ab directly to colloidal
gold, mainly because of the added cost (a similar concern of yours, I take
it!).

Are you up for a slightly different approach, which you can do with
commercially available reagents? If so, you could do the following:

1. On one side of the grid (maybe a fairly high hexagonal mesh, uncoated
nickel grid?) incubate with one of your mouse monoclonals.

2. Follow with biotinylated goat anti-mouse Ig-whatever (is it an IgG?).

3. Follow with streptavidin-gold (available in many different sizes).

4. Flip the grid over and repeat the labeling with the second monoclonal.
Only this time, use a different size of streptavidin-gold.

5. Double sided labeling works quite nicely for co-localizations, and it
eliminates a lot of cross-reactivity problems you'd have to fight if you
worked on only one surface of the section.

6. One thing to watch out for: If you add normal goat serum and maybe a
little TWEEN or other additives to the reagents, the grids will tend to sink
if they're incubated on drops of the reagents. Then you're labeling *both*
sides with the *same* reagents. This, of course, is to be avoided! You have
to keep the sides separated.

The best way to avoid this problem is to anchor the grid by its edge on a
piece of double-stick tape on a microscope slide. You can then use a
micropipette to deliver *very small* volumes (maybe 2-3 microliters) of each
reagent. To change reagents or to wash, simply add the reagent to one edge
of the grid and wick it off with filter paper
from the opposite edge.

Following the first gold incubation you can remove the grid and wash well in
a gentle stream or w/ repeated drops of buffer and finally water. Allow to
dry, flip the grid over, stick it down to the tape and do the second
run-through.

An added bonus of this procedure is a fairly good signal amplification. If
you conjugated your primary directly to the gold, you'd only have a one-step
reaction with minimal gold particles (probably 1) binding at each site.
Would that be enough to detect above background?

Unless, of course, you *want* the most specific localization possible...In
that case, forget everything I've just written!

Good luck!

Bob Chiovetti




From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Fri, 21 Mar 97 01:01:22 -0500
Subject: Flat embedding molds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Pat Hale wrote:
==================================================
I am currently trying to switch from Epon embedding media to one which
polymerizes by UV at lower temperatures. In order to orient my tissue
properly I would like to continue using flat embedding molds. My problem is
in making these molds "airtight" in order to prevent shrinkage/evaporation
while it polymerizes. Has anyone had this problem or found a solution to it?
==================================================
There is an alternative to the rigid polyethylene based (e. g. BEEM
manufactured) flat embedding molds, and that is the flexible and reusable UV
transparent silicone flat embedding molds. These silicone molds work just
fine and seem to have a bit higher transparency for the UV. One would
follow the method described by Bob Chiovetti, that is, to overfill the
cavities but instead of using Parafilm, just place the bottom of another one
of the UV transparent silicone molds on top. Capillary action quickly seals
out any entrapped air, creating the desired oxygen free environment.

UV transparent embedding molds come in varous sizes and shapes and are shown
on the SPI website given below. Similar, but not idential molds are also
offered by Ladd and possibly others.

Disclaimer: SPI Supplies manufactures UV transparent silicone embedding
molds and has a vested interest in seeing more persons using them!

Chuck


===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: http://www.2spi.com
############################
==================================================




From: WARRENJ1-at-cliffy.polaroid.com
Date: 3.20.97 5:09 PM
Subject: color 35mm neg scanner

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Polaroid Sprint Scan 35LE will scan your negatives in at
resolutions up to 1950 dpi -at- 10 bits per channel with an optical
density of 3.0 in under 60 seconds. It will scan 35 mm negatives or
slides(mounted or unmounted)as well as the SuperSize format. It has an
option called the PathScan Enabler that allow you to scan a standard
glass slide at the referenced resolution and color depth. It has a
list price of $995US.

John D. Warren
Area Sales Manager
Digital Products
Polaroid Corporation


______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I have hundreds of 35mm color negs which I wish to make proof sheets of.
I do not want to do anything with those images other than print proof sheets.
Can anyone suggest a good scanner for this job.

Thank you.
Ted Dunn




From: paulcd-at-bio.uva.nl (Paul)
Date: Fri, 21 Mar 1997 13:41:05 +0100
Subject: color 35mm neg scanner

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


POSITION AVAILABLE!!!!
At the Institute for Neurobiology, University of Amsterdam, The Netherlands
a post-doc position is available, for 3 years. The project concerns
plasticity in brain cells, in relation to epilepsy and hormonal influences.
It will make use of (fast) calcium imaging techniques with confocal
microscopy, in combination with patch clamp recording. We are looking for
someone with experience in confocal microscopy, preferably in brain tissue.
The position is available as of now. For more information:
wadman-at-bio.uva.nl or joels-at-bio.uva.nl or call 31205257641 / 31205257626.

-----------------
P.C.Diegenbach
dept. Biology, University of Amsterdam
Kruislaan 320 1098 SM Amsterdam, Netherlands
phone (31) 20 5257631 fax (31) 20 5257709
The following email address is mangled to prevent automated
unsolicited junk mail. Replace the '_AT_' with an '-at-':
email paulcd_AT_bio.uva.nl
-----------------






From: LUCY RU-SIU YIN :      lyin-at-oitunix.oit.umass.edu
Date: Fri, 21 Mar 1997 09:53:31 -0500 (EST)
Subject: Re: UV Embedding in Flat Molds

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Pat: I use UV polymerization of LR white or Unicryl at 4C for
immunolabelling studies. I use small aluminum weighing dish layered with
a sheet of Aclar film at the bottom. The embedding and tissue are placed
on top of this film. Cut a sheet of closely fit Aclar film to place over
it. All the weighing dishes are placed in a stainless tray which is sealed with
Saran Wrap. The aluminum will reflect the Uv and provide good even
polymerization. The embedding medium between the sheet will polymerize
while some excess media flow over and under the sheet will not. The
Aclar sheets are easily peeled off from the polymerized material. Sheets
of embedding material can be scanned under LM to chose for ideal
orientationed sample. Hope this will help you.

Licy Yin
Microscopy Center
U.Mass , Amherst,
MA 01003t
MiclOn Thu, 20 Mar 1997, pat
hales wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} I am currently trying to switch from Epon embedding media to one which
} polymerizes by UV at lower temperatures. In order to orient my tissue
} properly I would like to continue using flat embedding molds. My problem is
} in making these molds "airtight" in order to prevent shrinkage/evaporation
} while it polymerizes. Has anyone had this problem or found a solution to it?
}
} Pat Hales
} McGill University
} Dept. of Anatomy & Cell Biology
} hales-at-hippo.medcor.mcgill.ca
}




From: mme-at-map.com (barbara foster)
Date: Fri, 21 Mar 1997 09:59:59 -0800
Subject: slide films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear slide makers,

We have had superb results with Fuji's Velvia. It is a daylight film, so
you will need blue filtration when used with tungsten halogen sources
(82B)and it is slow (50ASA) but it produces exquisite color rendition and
depth. It is also a standard E-6, for easy development. Good luck...
and if you take an especially beautiful shot which you would like to
share, let us know. We are always looking for interesting images for our
short courses (with credit, of course).

Barbara Foster
Microscopy/Microscopy Education
53 Eton Street
Springfield, MA 01108
(413)746-6931 fax: (413)746-9311 email:mme-at-map.com




From: HILDEGARD CROWLEY :      hcrowley-at-odin.cair.du.edu
Date: Fri, 21 Mar 1997 08:24:10 -0700 (MST)
Subject: Re: Acrylics-oxygen-UV-polym.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Pat Hales asks about airtight seals for acrylic embedding medium.

Unicryl, an acrylic embedding medium, advertises that excluding oxygen
during UV polym is not
necessary. Unicryl is available from a number of vendors. (I have NO
interest in the commercial uses of Unicryl). Using an acrylic which
bypasses the entire problem of oxygen interference might be a good solution.
Bye,
Hildy




From: Nestor J. Zaluzec-Argonne Nat. Lab. :      ZALUZEC-at-aaem.amc.anl.gov
Date: Fri, 21 Mar 1997 11:28:39 -0600 (CST)
Subject: MidWest Microscopy & Microanalysis Society Workshop 4/11/97

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

MIDWEST MICROSCOPY AND MICROANALYSIS SOCIETY
affiliate of the Microscopy Society of America
and the Microbeam Analysis Society

presents

WORKSHOP AND SYMPOSIUM ON IN SITU HYBRIDIZATION

The program has been arranged by and will be held at Madison Area
Technical College (MATC), WI., Friday, April 11, 1996 from 8:00 - 5:00 p.m.
This meeting will consist of two concurrently held workshops and an
afternoon related symposium. The general emphasis will be in situ
hybridization. The two morning workshops will be on detection of
hybridization with radioactive and non radioactive probes.


The meeting has several purposes:

1. To draw the attention of the scientific community to emerging
developments in the practical and basic research aspects of
excitingnew fields.

2. To bring people together from diverse disciplines in order to
discuss how innovative techniques will be relevant to the future
direction of microscopy and microbeam analysis.

WORKSHOPS WILL BE HELD IN THE MORNING FROM 8:30 -12:30.

1. Radioactive in situ hybridization
Patricia L. Allen/Dr. Lyn Thet
Department of Pulmonary Medicine, School of Medicine
UW-Madison

2. Nonradioactive in situ hybridization
Boehringer Mannheim, Indianapolis, IN

You may choose to attend one workshop from above. Registration
information is enclosed. Register early because we will have to limit
the number of participants.




AFTERNOON SYMPOSIUM SCHEDULE
1:30 - 5:00 p.m.


Patricia Thomas, M.D.
Department of Pathology, School of Medicine, University of Iowa, Iowa
City, IA, "Diagnostic applications of fluorescence in situ
hybridization"

Kathrine Staskus, Ph.D.
Department of Microbiology, University of Minnesota, Minneapolis, MN,
"In situ hybridization detection of Simian Immunodeficiency virus
(SIV) proteins and nucleic acids"

Gary Lyons, Ph.D.
Department of Anatomy and Cellular and Molecular Biology Program,
School of Medicine, University of Wisconsin, Madison, "In situ
hybridization studies of murine embryonic skeletal muscle development"

Alan Smith, PhD.
Department of Horticulture, College of Agriculture, University of
Minnesota, Minneapolis, MN, "Localization of floral specific gene
products during pollen development"

Gwen V. Childs, Ph.D.
Vice Chair, Department of Anatomy and Neuroscience, Program Director.
Cell Biology Graduate Program, University of Texas Medical Branch,
Galveston, TX, "Regulation of multipotential cells in the pituitary by
neuroendrocrine peptide:an in situ hybridization and
immunocytochemical study"

Kevin Roth, M.D., Ph.D
Department of Pathology, Washington University, St. Louis, MO.
"Application of tyromide-signal amplification to in situ and
immunocytochemistry"


WORKSHOP AND SYMPOSIUM ON IN SITU HYBRIDIZATION

GENERAL INFORMATION SHEET

DATE: Friday, April 11, 1997

LOCATION: Madison Area Technical College (Truax Campus)
Electron Microscopy Program (Check-in, Rm375A)
3550 Anderson St., Madison, WI. 53704

Workshops and Symposium locations will be provided
at the Check-in room (Rm375A)

DIRECTIONS
PARKING: A map is provided on request. Please park in the large lot on
the side of the campus building.


MEALS: Refreshments will be provided in the morning and
for afternoon break.
Lunch is a sandwich buffet, and the cost is
$4.00/person. Following the afternoon symposium
there will be a wine, soft drink, and cheese
reception provided at no cost to participants.
An Italian dinner buffet is being organized for
after the wine and cheese reception. The cost is
$7.00/person.

REGISTRATION: If you plan to attend either one of the workshops or
the symposium you must register by phone or EMAIL to one of the
contacts listed below. Registration is free to all Society members,
student, regular, and corporate. If you are not a member of the
Society, registration cost will be the price of membership. (Regular =
$10, Student = $5). When you register you must provide us with the
information listed below. Name tags and Fees for registration and
meals will be collected at the Check-in site.

DEADLINE DATE FOR REGISTRATION IS APRIL 8.

1. Name
2. Affiliation
3. Phone #
4. Choice of one Workshop
5. Are you attending the Afternoon Symposium?
6. Do you want to participate in the Lunch or Dinner?

REGISTRATION CONTACTS:

James P. DiOrio
(847) 270-4676
Fax: (847) 270-4414
EMAIL: diorio-at-baxter.com

Joanne M. Crudele
(847) 734-3712
Fax: (847) 734-3686
EMAIL: Joanne.Crudele-at-Unilever. Com

If you need information on hotels, contact Michael Kostrna
(608)246-6762 or Kenneth Muse (608)243-4309.




From: Ming Chen :      mingchen-at-gpu.srv.ualberta.ca
Date: Fri, 21 Mar 1997 10:36:22 -0700 (MST)
Subject: Re: preparation/coating of beeswax

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Thu, 20 Mar 1997, Ginger Baker wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello all,
}
} I have a PhD student from Brazil who is studying Catolaccus
} grandis. He places a female on top of a beeswax cell which contains
} boll weevil larva. The female deposits an egg on the larva and 2
} weeks later, it matures and burrows its way out of the beeswax
} creating a hole. He would like to study and measure these holes.
}
} However, from past experience I know that beeswax melts when placed
} in my Au/Pd sputter coater. Is there any way to coat the beeswax and
} examine them with an SEM? Or should he try another route?
}
} Thank you in advance,

} Email: lizard-at-okway.okstate.edu
}
Do you have cooling system on the sputter coater ? If your answer is
yes, let the beeswas sit on the cooling stage for 20 min then coat it
for 5-10 sec and pause for about 20 sec. Repeat it for 3-4 cycles. My
experience for fibric material the above tricks would work fine. Also you
may lower your sample far from the gold target to avoid heat.

Hope this will help.


***********************************************
* Ming H. Chen, PhD *
* Medicine/Dentistry Electron Microscopy Unit *
* University Of Alberta. *
* Edmonton, Alberta, Canada *
* *
* Visit My Page At: *
* http://www.ualberta.ca/~mingchen *
***********************************************








From: sbarlow-at-sunstroke.sdsu.edu (Steve Barlow)
Date: Fri, 21 Mar 1997 09:49:20 -0800
Subject: Meeting/Call for Papers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


5th Annual
CSU Microscopy Colloquium

Saturday, April 26, 1997

San Jose State University

With a Special Lecture and Optional Demonstration Workshop
on in situ Hybridization on Sunday, April 27
presented by Dr. L=E1szl=F3 K=F6m=FCves of UC San Francisco

Business meeting for CSU delegates, Friday, April 25, 1997 at 4:00 PM in
Duncan Hall Rm. 249

Proceedings will be published in Microscopy Research and Technique

The Colloquium provides a forum for the exchange of research results and
experiences in all fields of light and electron microscopy in the
biological, geological, and materials sciences. Participants include
students and scientists from academia and industry.

Both platform and poster presentations are invited.
Student presentations are strongly encouraged.

Registration Fees $35 Regular, $10 Student, $50 Vendor by March 15; $25
Optional Workshop

Late Registration $45 Regular, $10 Student, $60 Vendor, after March 15;
$25 Optional Workshop

Make checks payable to San Jose State University Foundation

Send to: Dr. David K. Bruck
CSU Microscopy Colloquium
Department of Biological Sciences
San Jose State University
San Jose, CA 95192-0100

=46or additional information, flyers, and abstract forms, contact
David Bruck at phone: (408) 924-4837 or fax (408) 924-4840
bruck-at-biomail.sjsu.edu

Program

=46riday, April 25

4:00 PM Business meeting for CSU delegates in Duncan Hall Rm. 249, San Jose
State University

Saturday, April 26

8:00 AM Registration, coffee, and breakfast snacks in lobby of Duncan Hall,
San Jose State University
9:00 AM Opening remarks in Duncan Hall Rm. 135
9:15 AM Platform presentations
10:30 AM Coffee break
11:00 AM Platform presentations
12:00 PM Buffet lunch
1:30 PM Platform presentations
2:30 PM Coffee break
3:00 PM Platform presentations
4:00 PM Lecture on in situ hybridization by Dr. Laszlo Komuves
5:00 PM Closing Remarks and Poster session in Duncan Hall Rm. 250
6:30 PM Banquet

Sunday, April 27

9:00 AM Demonstration and workshop on in situ hybridization conducted by
Dr. Laszlo Komuves in Duncan Hall Rm. 344

Housing

Best Western Inn
455 S. Second St.
San Jose, CA 95113
phone (800) 528-1234
or (408) 298-3500
singles $52, doubles $62

Please make your own reservations for housing at one of the phone numbers
above. Specify that you will be attending the Microscopy Colloquium to
receive the conference discount. A map and other hotels in the vicinity
will be included in the packet of information mailed to you following
receipt of your registration information.

Transportation

=46ree parking on campus is available in the 7th St. Garage on the corner of
7th St. and San Salvador St. A permit will be mailed to you following
receipt of your registration form.

Public transportation from the San Jose International Airport to downtown
San Jose is limited to cabs or the Light Rail system. For the latter, a
free shuttle bus (labeled Metro) runs every 10-15 minutes from the airport
to the Metro Light Rail Station. For $1.10, the train runs downtown to the
Paseo de San Antonio Station, about 3 blocks from the university and 2
blocks from the Best Western Inn.


Registration

Please complete the registration form below whether or not you will be
presenting a paper or poster.

Registration Fees
$35 regular, $10 student, $50 vendor

After March 15
$45 regular, $10 student, $60 vendor

_____________________________
Name

_____________________________
Address

_____________________________
City State Zip

_____________________________
Phone

_____________________________
e-mail

Workshop Registration
(for an additional fee of $25)
0 Workshop

Note that the workshop's occurrence will depend on an adequate number of
registrants.

Meal Preference
0 Chicken 0 Fish 0 Vegetarian

Presentation
0 Platform 0 Poster 0 None

Information and Mailing

Make registration checks payable to the San Jose State University
=46oundation. Mail registration forms, abstract forms, and checks to

Dr. David K. Bruck
CSU Microscopy Colloquium
Department of Biological Sciences
San Jose State University
1 Washington Square
San Jose, CA 95192-0100

(408) 924-4837
bruck-at-biomail.sjsu.edu

Platform and Poster Presentations

Papers discussing any aspect of microscopy in cell, developmental, and
structural biology or in the material or geological sciences are welcome.
Microscopy techniques' papers are particularly encouraged.

Platform presentations (15 min.) will be held Saturday, April 26. Posters
will be displayed all day Saturday, with an afternoon session designated
for viewing with authors present.

An Abstract Guidelines form is included for those presenting papers or
posters. The abstracts will be compiled into the "Proceedings of the Fifth
Annual CSU Microscopy Colloquium" and will be published in Microscopy
Research and Technique.




---------------------------------------------------------------------

Dr. Steven Barlow
EM Facility/Biology Department
5500 Campanile Drive
San Diego CA 92182-4614
phone: (619)594-4523
fax: (619) 594-5676
email: sbarlow-at-sunstroke.sdsu.edu
website: http://www.sci.sdsu.edu/EM_Facility






From: gradice-at-richmond.edu (Gary Radice)
Date: Fri, 21 Mar 1997 15:55:22 -0500
Subject: Aclar film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Several of you have recommending using Aclar film in your UV polymerization
protocols. What is Aclar film, and where do you purchase it?

Gary Radice, Associate Professor gradice-at-richmond.edu
Department of Biology 804-289-8107 (voice)
University of Richmond VA 23173 804-289-8233 (FAX)






From: John J. Lemasters :      lemaster-at-med.unc.edu
Date: Fri, 21 Mar 1997 15:59:44 -0500 (EST)
Subject: Carolina Workshop on LIGHT MICROSCOPY

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


PLEASE NOTE CORRECTED E-MAIL ADDRESS FOR DR. LITAKER

-------------------------------------------------------------------
COURSE ANNOUNCEMENT

Carolina Workshop on: LIGHT MICROSCOPY FOR THE BIOMEDICAL SCIENCES
June 1-6, 1997
University of North Carolina at Chapel Hill

Instructors: John J. Lemasters
Edward D. Salmon
Brian Herman
-------------------------------------------------------------------
LIGHT MICROSCOPY FOR THE BIOMEDICAL SCIENCES is an introduction
to applications of light microscopy. Students will have
opportunities for extensive hands-on experience with
state-of-the-art equipment for optical imaging, digital imaging
processing, fluorescence microscopy and confocal microscopy guided
by experienced academic and commercial staff. The course is
divided into three major sections with lectures and laboratory
exercises on: 1) geometric and wave optics of image formation,
microscope alignment, phase contrast and reflection interference
contrast microscopy; 2) video imaging, including contrast
enhancement by analog and digital image processing, fluorescence
microscopy, image detectors, fluorescent probes, ion imaging,
and green fluorescent protein; and 3) laser scanning confocal
microscopy emphasizing live cell imaging and 3-dimensional image
reconstruction. Students are encouraged to bring their own
specimens for analysis. A commercial staff representing leading
microscopic manufacturers will make available for student use the
latest and most advanced instrumentation for light microscopy, image
detection and computerized image analysis.

Tuition is $950.

-------------------------------------------------------------------
APPLICATION FORM
Carolina Workshop on LIGHT MICROSCOPY FOR THE BIOMEDICAL SCIENCES


Position:

Address:



Telephone:

Fax:


Please return this form along with a brief letter describing your
research interests and a curriculum vitae. Applicants
should contact the program as soon as possible. Full
consideration will be given to applications received by
April 18, 1997.

Send application to:
Dr. Wayne Litaker, Director of Workshops
University of North Carolina at Chapel Hill
Program in Molecular Biology &
Biotechnology
CB# 7100, 442 Taylor Hall
Chapel Hill, North Carolina 27599-7100
Tel: (919) 966-1730
Fax: (919) 966-6821
e-mail: litaker-at-med.unc.edu
-------------------------------------------------------------------
{End of Announcement}




From: Eric :      earosen-at-pop.goodnet.com
Date: Fri, 21 Mar 1997 16:59:07 -0800
Subject: DIC alignmetn on Nikon microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have one single question for the wealth of knowledge out there in
the world of Microscopy....

On a Nikon Diaphot inverted with a universal Con.A Achr-Apl 1.4
condenser using a 10X objective with DIC do you have to oil the
condenser????

Also how do you do the alignment properly, since noone in the lab I
work knows... I think I have everything correct but when I twist
polarizer by the objective lens I do not get the rainbow of colors or
even the black background when the polarizers are ligned up....


Any Suggestions
\\|//
(o o)
~~~~~~~~~~~~oOOo~(_)~oOOo~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

{ { {This message is made of 100% recycled electrons} } }

Cheers ;o) :o) %o)
Eric {Mesa Arizona}
http://www.goodnet.com/~earosen (Note the tilde before earosen)




From: Ronald LHerault :      lherault-at-bu.edu
Date: Fri, 21 Mar 1997 20:04:48 -0500 (EST)
Subject: Prep of bees wax.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

An alternative may be replication using Reprosil, a Polyvinyl Siloxane
from Caulk. It makes a negative which can be transfered to a positive by
using Spurr low viscosity embedding medium. Detail is there albeit a
little softened.

Ron





From: Sun Qi-an :      sunqa-at-mail.jlu.edu.cn
Date: Sat, 22 Mar 1997 17:06:25 -0800
Subject: Sun's Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I am a student now using Hitachi-8100 TEM to study submicrometer Cu-Zn
alloy particals. The particals are mostly six-fold symmetry plates. When
the beam focused on a partical at a high magnification, some dark
stripes appear in the partical's bright-field image and change their
shapes until a snow-flake-like pattern is formed. The pattern comprised
of six center-crossed dark stripes. Select-area diffracton shows a
single crystal diffracton pattern with six-fold symmetry. But dark-field
image shows that one single diff spot is not coming from the whole
partical but from ONE corresponding dark stripe.

My questions are: 1)Do anyone out there ever find similar phenomenon?
2)How can I know the temperature of the particals
under the brightening of the electron
beam? I think temperature is a reason for the
formation of the snow-flake-like pattern.

I hope I have described my question clear enough for you to reply.
Thanks!

Sun Haiping
National Lab of Superhard Materials
Jilin University
Changchun 130023, P.R.China
sunqa-at-mail.jlu.edu.cn




From: Jim Darley :      jim-at-proscitech.com.au
Date: Sat, 22 Mar 1997 22:27:36 +1000
Subject: Re: Aclar film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Aclar is a thin plastic sheet material suitable for cell culture and may be
sectioned. A full description is in our catalogue on page L7.
We distribute Aclar as does Ted Pella (Pelco) in the USA.
Jim Darley

ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 77 740 370 Fax: +61 77 892 313
Great microscopy catalogue, 300+ Links, MSDS
************************ http://www.proscitech.com.au

} Several of you have recommending using Aclar film in your UV
polymerization
} protocols. What is Aclar film, and where do you purchase it?
}
} Gary Radice, Associate Professor gradice-at-richmond.edu
} Department of Biology 804-289-8107 (voice)
} University of Richmond VA 23173 804-289-8233 (FAX)
}
}




From: H. ADAMS :      hadams-at-nmsu.edu
Date: Sat, 22 Mar 1997 09:40:42 -0700 (MST)
Subject: Lectin staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello, I am having no luck attempting to stain the glycocalyx? of a
species of sulfate-reducing bacteria using ConA-conjugated gold (5nm).
Using 4%PF+0.1%Glute in Cac buffer fixation followed by LRW embedding I
have tried 1:10 to 1:100 in P04 buffer pH 7.2. I have also tried ConA-gold
in Tris-saline with and without 1mM CaCl2 and 1mM MgCl2 at pH 6.9,7.2
and 8.0. diluted 1:10-1:100. Next I tried taking a fresh culture and
washing several times in Tris-saline followed by 1:10 ConA-gold,then
fixing and embedding. My incubations have been around 1 hr at room temp.
Nary a gold particle do Isee under any of the conditions mentioned.
Apparently, using fluorescently labelled ConA these little buggers lightup
like the Hale-Bopp comet, so Iam at a lost. Can anybody out there give me
some suggestions. I also have some material that I fixed in 4% PF alone,
embedded in LRW that I have not done anything yet with.
Thanks in advance

Hank Adams
Electron Microscopy Laboratory
New Mexico State University
Las Cruces, NM 88003

http://www.nmsu.edu/Research/artsci/public_html/eml/

505-6463600







From: David Chase :      dchase-at-gwi.net \
Date: Sat, 22 Mar 1997 15:12:49 -0500
Subject: A O scopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have an American Optical Micro-Star Illuminator
model 1872.
It has 10XB&L WF eyepieces and a third place for
camera/eyepiece.
The primary voltage is 115 and secondary 3 to 6
volts using GE lamp 1974
It has no objectives, but spaces for 4.
It has a large mechanical base.
It is illuminated thru the lenses.

I know nothing else about the scope except it was
used in a lab at
Fairchild semiconductor in South.

Iam looking for some good quality/inexpensive
objectives for amatuer use
and any manuals/info on using this scope.

If you know anything of this scope or the objectives
that I need, please
e-mail me

Thanks so much for your time

David Chase
dchase-at-gwi.net

ps I had researched this all once before but a fatal
HD crash destroyed
all my data.






From: H. ADAMS :      hadams-at-nmsu.edu
Date: Sat, 22 Mar 1997 15:20:12 -0500
Subject: Lectin staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello, I am having no luck attempting to stain the glycocalyx? of a
species of sulfate-reducing bacteria using ConA-conjugated gold (5nm).
Using 4%PF+0.1%Glute in Cac buffer fixation followed by LRW embedding I
have tried 1:10 to 1:100 in P04 buffer pH 7.2. I have also tried ConA-gold
in Tris-saline with and without 1mM CaCl2 and 1mM MgCl2 at pH 6.9,7.2
and 8.0. diluted 1:10-1:100. Next I tried taking a fresh culture and
washing several times in Tris-saline followed by 1:10 ConA-gold. My
incubations have been around 1 hr at room temp. Nary a gold particle do I
see under any of the conditions mentioned. Apparently, using fluorescently
labelled ConA these little buggers light up like the Hale-Bopp comet, so I
am at a lost. Can anybody out there give me some suggestions. I also have
some material fixed in 4% PF alone, embedded in LRW that I have not done
anything yet with.
Thanks in advance,

Hank Adams
Electron Microscopy Laboratory
New Mexico State University
Las Cruces, NM 88003

http://www.nmsu.edu/Research/artsci/public_html/eml/

505-6463600






From: Paul Webster :      paul.webster-at-Yale.edu
Date: 22 Mar 1997 18:39:10 -0500
Subject: Re: Lectin Staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hank Adams writes:
"Hello, I am having no luck attempting to stain the glycocalyx? of a
species of sulfate-reducing bacteria using ConA-conjugated gold (5nm)."
(trimmed).

Dear Hank,
Your problem may lie in the ConA-conjugated gold. Often, these probes very
seldom store well. If most of the protein has dissociated from the gold
particles, you will have a large amount of very specific ConA binding but no
gold to show you it is there. It is easy to test at the LM level by probing
sections treated with ConA-god with anti-ConA and fluorescent antibody.

Once again, the best visualization method for lectins (re:ricin) may be to use
antibodies instead of the directly conjugated gold probes. If you prefer to use
ConA-gold then I recommend that you make your own. A slightly less satisfactory
approach would be for you to centrifuge down your ConA-gold (be careful not to
spin it so hard that is aggregates into the bottom of the tube) and resuspend it
in fresh PBS. This will remove most of the free protein.

As an aside and in reference to a previous posting, streptavidin-gold also
suffers from this storage problem. When using the probe at the LM level, with
silver enhancement, or with multiple antibody layers the loss of labeling
efficiency is not so noticeable. However, attempting to visualize biotinylated
antibodies directly with streptavidin-gold will only work well for the first
week after the probe has been made. For situations where there are large
numbers of bound antibody, this effect may not be noticeable, but for situations
where there are only low numbers of antigens, it is very obvious.

Making colloidal gold and coupling it to proteins or antibodies is easy and
anyone with even limited experience can do it.

Check out my WWW "gold page" to see how easily it can be done. Be warned, if
you want to do this you usually need large supplies of protein (a point anyone
coupling antibodies should be aware of).

Paul Webster
Center for Cell Imaging
Yale School of Medicine
http://info.med.yale.edu/cellimg





From: Eric :      earosen-at-pop.goodnet.com
Date: Sat, 22 Mar 1997 20:46:17 -0800
Subject: Re: Lectin staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {199703230345.UAA07299-at-mailque.goodnet.com}
Comments: Authenticated sender is {earosen-at-pop.goodnet.com}

You may also want to treat the sections with a very low concentration
of sodium periodate which will destroy the section if you are not
careful.. This will expose some of the bnding sites fo the Con-A.. or
on the other hand the Lctins are too old, and the gold has fallen off
of the lectin which occured to me also when I was using gold labeled
lectins that were stored in the fridge........ So I got very
scattered labeling or no labelling at all but when used in teh LM
with Cy3 they light up like a chrsitmas tree....

lAlso you may want to try a lower comcentration of fixatives in
Formaldehyde and Glut.


---------------------------------------------------------------------
--- The Microscopy ListServer -- Sponsor: The Microscopy Society of
America To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
----------------------------------------------------------------------
-.

Hello, I am having no luck attempting to stain the glycocalyx? of a
species of sulfate-reducing bacteria using ConA-conjugated gold (5nm).
Using 4%PF+0.1%Glute in Cac buffer fixation followed by LRW embedding
I have tried 1:10 to 1:100 in P04 buffer pH 7.2. I have also tried
ConA-gold in Tris-saline with and without 1mM CaCl2 and 1mM MgCl2 at
pH 6.9,7.2 and 8.0. diluted 1:10-1:100. Next I tried taking a fresh
culture and washing several times in Tris-saline followed by 1:10
ConA-gold. My incubations have been around 1 hr at room temp. Nary a
gold particle do I see under any of the conditions mentioned.
Apparently, using fluorescently labelled ConA these little buggers
light up like the Hale-Bopp comet, so I am at a lost. Can anybody out
there give me some suggestions. I also have some material fixed in 4%
PF alone, embedded in LRW that I have not done anything yet with.
Thanks in advance,

Hank Adams
Electron Microscopy Laboratory
New Mexico State University
Las Cruces, NM 88003

http://www.nmsu.edu/Research/artsci/public_html/eml/

505-6463600



\\|//
(o o)
~~~~~~~~~~~~oOOo~(_)~oOOo~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

{ { {This message is made of 100% recycled electrons} } }

Cheers ;o) :o) %o)
Eric {Mesa Arizona}
http://www.goodnet.com/~earosen (Note the tilde before earosen)




From: RA :      ralpha-at-softcom.net
Date: Sat, 22 Mar 1997 20:57:53 -0800
Subject: Looking for Microscopist newbies newsgroups..

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Can any one tell me of a Microscope news group taylored toward the
newbie into this hobby? Also, are there any good magazines on the market
specializing in MIcroscopes and the Microscope hobby?

Thanks for any information.

Ralph




From: Sara Miller :      saram-at-acpub.duke.edu
Date: Sun, 23 Mar 1997 14:18:10 -0500 (EST)
Subject: Re: Lectin staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Sat, 22 Mar 1997, H. ADAMS wrote:

} Date: Sat, 22 Mar 1997 15:20:12 -0500
} From: H. ADAMS {hadams-at-nmsu.edu}
} To: microscopy-at-Sparc5.Microscopy.Com
} Subject: Lectin staining
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello, I am having no luck attempting to stain the glycocalyx? of a
} species of sulfate-reducing bacteria using ConA-conjugated gold (5nm).
} Using 4%PF+0.1%Glute in Cac buffer fixation followed by LRW embedding I
} have tried 1:10 to 1:100 in P04 buffer pH 7.2. I have also tried ConA-gold
} in Tris-saline with and without 1mM CaCl2 and 1mM MgCl2 at pH 6.9,7.2
} and 8.0. diluted 1:10-1:100. Next I tried taking a fresh culture and
} washing several times in Tris-saline followed by 1:10 ConA-gold. My
} incubations have been around 1 hr at room temp. Nary a gold particle do I
} see under any of the conditions mentioned. Apparently, using fluorescently
} labelled ConA these little buggers light up like the Hale-Bopp comet, so I
} am at a lost. Can anybody out there give me some suggestions. I also have
} some material fixed in 4% PF alone, embedded in LRW that I have not done
} anything yet with.
} Thanks in advance,
}
} Hank Adams
} Electron Microscopy Laboratory
} New Mexico State University
} Las Cruces, NM 88003
}
} http://www.nmsu.edu/Research/artsci/public_html/eml/
}
} 505-6463600
}
Since the glycocalyx is on the outside of the bacteria and you don't have
to worry about sectioning to expose the conA to the glycocalyx, why don't
you stain with con A-Au, wash thoroughly (centrifugations), and then embed
normally?

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735





From: Sara Miller :      saram-at-acpub.duke.edu
Date: Sun, 23 Mar 1997 14:31:54 -0500 (EST)
Subject: Re: Lectin Staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On 22 Mar 1997, Paul Webster wrote:

} Date: 22 Mar 1997 18:39:10 -0500
} From: Paul Webster {paul.webster-at-Yale.edu}
} To: "Microscopy -at-MSA.Microscopy.Com" {Microscopy-at-Sparc5.Microscopy.Com}
} Subject: Re: Lectin Staining
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hank Adams writes:
} "Hello, I am having no luck attempting to stain the glycocalyx? of a
} species of sulfate-reducing bacteria using ConA-conjugated gold (5nm)."
} (trimmed).
}
} Dear Hank,
} Your problem may lie in the ConA-conjugated gold. Often, these probes very
} seldom store well. If most of the protein has dissociated from the gold
} particles, you will have a large amount of very specific ConA binding but no
} gold to show you it is there. It is easy to test at the LM level by probing
} sections treated with ConA-god with anti-ConA and fluorescent antibody.
}
} Once again, the best visualization method for lectins (re:ricin) may be to use
} antibodies instead of the directly conjugated gold probes. If you prefer to use
} ConA-gold then I recommend that you make your own. A slightly less satisfactory
} approach would be for you to centrifuge down your ConA-gold (be careful not to
} spin it so hard that is aggregates into the bottom of the tube) and resuspend it
} in fresh PBS. This will remove most of the free protein.
}
} As an aside and in reference to a previous posting, streptavidin-gold also
} suffers from this storage problem. When using the probe at the LM level, with
} silver enhancement, or with multiple antibody layers the loss of labeling
} efficiency is not so noticeable. However, attempting to visualize biotinylated
} antibodies directly with streptavidin-gold will only work well for the first
} week after the probe has been made. For situations where there are large
} numbers of bound antibody, this effect may not be noticeable, but for situations
} where there are only low numbers of antigens, it is very obvious.
}
} Making colloidal gold and coupling it to proteins or antibodies is easy and
} anyone with even limited experience can do it.
}
} Check out my WWW "gold page" to see how easily it can be done. Be warned, if
} you want to do this you usually need large supplies of protein (a point anyone
} coupling antibodies should be aware of).
}
} Paul Webster
} Center for Cell Imaging
} Yale School of Medicine
} http://info.med.yale.edu/cellimg
}
Also, if you're going to spin out your con A-Au from the uncongugated con
A, keep the pH high (probably above 8-8.2, or whatever pH the original
stuff was shipped in). Otherwise, you might have one big blob at the
bottom of your tube. When we suspect unconjugated protein, we dilute the
antibody-Au to the concentration we want to use with buffer at a pH above
the pI (pH } 8), and then spin in a Beckman Airfuge 5 min at 30 lb
(~100,000 g), discard the sup and resuspend the pellet in the same vol(pH 7)
as we started with. Haven't tried con A; I assume it might work the same
way???

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735





From: Sara Miller :      saram-at-acpub.duke.edu
Date: Sun, 23 Mar 1997 14:35:24 -0500 (EST)
Subject: Re: Lectin staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Sat, 22 Mar 1997, Eric wrote:

} Date: Sat, 22 Mar 1997 20:46:17 -0800
} From: Eric {earosen-at-pop.goodnet.com}
} To: microscopy-at-Sparc5.Microscopy.Com
} Subject: Re: Lectin staining
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} You may also want to treat the sections with a very low concentration
} of sodium periodate which will destroy the section if you are not
} careful.. This will expose some of the bnding sites fo the Con-A.. or
} on the other hand the Lctins are too old, and the gold has fallen off
} of the lectin which occured to me also when I was using gold labeled
} lectins that were stored in the fridge........ So I got very
} scattered labeling or no labelling at all but when used in teh LM
} with Cy3 they light up like a chrsitmas tree....
}
} lAlso you may want to try a lower comcentration of fixatives in
} Formaldehyde and Glut.
}
}
} ---------------------------------------------------------------------
} --- The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} ----------------------------------------------------------------------
} -.
}
} Hello, I am having no luck attempting to stain the glycocalyx? of a
} species of sulfate-reducing bacteria using ConA-conjugated gold (5nm).
} Using 4%PF+0.1%Glute in Cac buffer fixation followed by LRW embedding
} I have tried 1:10 to 1:100 in P04 buffer pH 7.2. I have also tried
} ConA-gold in Tris-saline with and without 1mM CaCl2 and 1mM MgCl2 at
} pH 6.9,7.2 and 8.0. diluted 1:10-1:100. Next I tried taking a fresh
} culture and washing several times in Tris-saline followed by 1:10
} ConA-gold. My incubations have been around 1 hr at room temp. Nary a
} gold particle do I see under any of the conditions mentioned.
} Apparently, using fluorescently labelled ConA these little buggers
} light up like the Hale-Bopp comet, so I am at a lost. Can anybody out
} there give me some suggestions. I also have some material fixed in 4%
} PF alone, embedded in LRW that I have not done anything yet with.
} Thanks in advance,
}
} Hank Adams
} Electron Microscopy Laboratory
} New Mexico State University
} Las Cruces, NM 88003
}
} http://www.nmsu.edu/Research/artsci/public_html/eml/
}
} 505-6463600
}
}
}
} \\|//
} (o o)
} ~~~~~~~~~~~~oOOo~(_)~oOOo~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
}
} { { {This message is made of 100% recycled electrons} } }
}
} Cheers ;o) :o) %o)
} Eric {Mesa Arizona}
} http://www.goodnet.com/~earosen (Note the tilde before earosen)
}


Etching agents are not usually required for porous acrylics.

If you want to test your fixation, try various ones and then fluorescent
labeling--saves a lot of work doing it at the EM level.

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735





From: Yi Huang :      yi_huang-at-qmgate.anl.gov
Date: 23 Mar 1997 17:30:40 -0600
Subject: Re: Sun's Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
sunqa-at-mail.jlu.edu.cn
X-Mailer: Mail*Link SMTP-QM 4.0.0



From: Yi Huang :      yi_huang-at-qmgate.anl.gov
Date: 23 Mar 1997 17:30:40 -0600
Subject: Re: Sun's Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Reply to: RE} Sun's Question

What you saw probably were bend extinction contour. (See Hersch et al's book
"Electron microscopy of thin crystals", page 203) The change of the pattern
may be caused by beam heating.

--------------------------------------
You wrote:

I am a student now using Hitachi-8100 TEM to study submicrometer Cu-Zn
alloy particals. The particals are mostly six-fold symmetry plates. When
the beam focused on a partical at a high magnification, some dark
stripes appear in the partical's bright-field image and change their
shapes until a snow-flake-like pattern is formed. The pattern comprised
of six center-crossed dark stripes. Select-area diffracton shows a
single crystal diffracton pattern with six-fold symmetry. But dark-field
image shows that one single diff spot is not coming from the whole
partical but from ONE corresponding dark stripe.

My questions are: 1)Do anyone out there ever find similar phenomenon?
2)How can I know the temperature of the particals
under the brightening of the electron
beam? I think temperature is a reason for the
formation of the snow-flake-like pattern.







From: Michael OKeefe :      Michael_OKeefe-at-macmail.lbl.gov
Date: 23 Mar 1997 16:27:04 -0700
Subject: Position Available at NCEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-ID: {n1352987827.40745-at-macmail.lbl.gov}

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Subject: Time:4:09 PM
OFFICE MEMO Position Available at NCEM Date:3/23/97

Staff Materials Scientist

Materials Sciences Division
National Center for Electron Microscopy (NCEM)
Job MSD/5001

Essential --
The NCEM is a national user facility with several state-of-the-art
electron microscopes and advanced image analysis and specimen
preparation facilities, dedicated to the electron-optical
microcharacterization of materials.
The Center has an opening for a Staff Scientist to lead its research
effort in high-resolution imaging of materials.
Conduct original research into the development of high-resolution
techniques and their application to significant materials problems.
Responsible for the operation and development of several high-
resolution microscopes, including a new 300kV field emission
instrument and the 1MeV ARM.
Lead a program to extend instrument resolution toward the 1A information
limit.

As an essential member of a national user facility, the candidate
will have the opportunity to collaborate with a broad range of
investigators from the national and international scientific community.
The position also involves close collaboration with other staff members
on the development of computer analysis, and control and supervision of
postdoctoral and technical staff.

The candidate will be able to contribute substantially to the successful
operation and future development of the facility.

QUALIFICATIONS:
Essential --
Outstanding track record in development and application of quantitative
high-resolution transmission electron microscopy.
Extensive practical experience with advanced high-resolution imaging
and analysis techniques, such as series reconstruction, holography, and
image interpretation.
Demonstrated ability to initiate collaborations and carry out high-
quality research, using the unique facilities of the NCEM.
Marginal --
Ph.D. in physical sciences.

POSTING DATE: March 6, 1997.
CLOSING DATE: Open until filled.


MORE INFORMATION:

1. See -- http://www.lbl.gov/LBL-Documents/CJOs/
2. Go to SCIENTIFIC
3. Go to Job MSD/5001

APPLICATION INSTRUCTIONS
(http://www.lbl.gov/LBL-Documents/CJOs/)
You may forward your resume or curriculum vitae to:
Berkeley Lab Staffing Office
One Cyclotron Road, MS 938A
Berkeley, CA 94720

Refer to job number MSD/5001 in your cover letter.

E-Mail to: employment-at-LBL.gov

1. Send as plain text (ASCII) in the body of your message.
(Attachments or encoded files cannot be read
by our automated applicant processing system.)
2. Reference the job number(s) in the SUBJECT of your message.
3. You will not receive an electronic response but will receive
a letter of confirmation via the U.S. Postal Service.








From: Jussi.Liipo-at-outokumpu.fi (Jussi Liipo)
Date: Mon, 24 Mar 1997 08:17:29 +0200
Subject: LM - point counter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0



I wonder if anayone know where I can purchase a new point counter
(e.g. J. Swift) and accessories required for counting mineral
grains.

Regards,
Jussi




From: Ian MacLaren :      I.MacLaren-at-BHAM.AC.UK
Date: Mon, 24 Mar 1997 10:16:15 +0000
Subject: Re: Sun's Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Your labelling problem may simply be that you are using resin sections which
are usually not permeable to gold labels (ie you only present a very small
edge of the bacteria in e.m. whereas I assume the light microscopy that you
mention was done on whole cells which present a much larger surface to the
fluorescent label. One test might be to try and use the fluorescent label on
LR White sections and examine that under the fluorescent light microscope.

Malcolm Haswell
University of Sunderland
UK
----------

Dear Sun,
As someone else stated, the dark lines you saw in bright field are bend
contours. You saw three crossed bend contours, each which corresponds to a
particular diffraction condition. Thus where all three bend contours
cross, all three diffraction conditions are operating simultaneously and
you are looking down the zone axis of the crystal. When you work in dark
field you will be imaging with only one of the diffraction spots so you
will only see one bend contour (which will now be bright).
As for why the pattern changes on observation, this is quite common and
you will find that many things deform or tilt slightly under the heat of
the beam. I don't know about your system but in some systems, the heating
can induce crystallisation which would obviously change the contrast
pattern.
Lastly, I can't think of a simple way of measuring the temperature of your
particles.

Yours sincerely


++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Ian MacLaren, Tel: (44) (0) 121 414 3447
IRC in Materials for FAX: (44) (0) 121 414 3441
High Performance Applications, email: I.MacLaren-at-bham.ac.uk
The University of Birmingham, http://web.bham.ac.uk/I.MacLaren/
Birmingham B15 2TT,
England.
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++






From: RA
Date: 23 March 1997 06:53
Subject: Looking for Microscopist newbies newsg

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


If you have full World Wide Web access you could try using a search engine
such as Lycos or Yahoo (eg microscope, microscopy, amateur, hobby). It
should turn up some useful web pages which could lead to further links.

One really excellent on-line organization & magazine is:
Microscopy-uk
http://www.microscopy-uk.org.uk/#top
which has the online magazine 'Micscape'. It has had excellent articles on
everything from hand lenses to electron microscopes and makes few
assumptions that it's readership are pro microscopists. There are also lots
of links.

Other places to try would be the home pages of the Microscopy Society of
America (US) and Royal Microscopy Society (UK) and again look for useful
links.

I hope this helps

Malcolm Haswell
University of Sunderland
UK

----------

Can any one tell me of a Microscope news group taylored toward the newbie
into this hobby? Also, are there any good magazines on the market
specializing in MIcroscopes and the Microscope hobby?

Thanks for any information.

Ralph





From: Sun Haiping :      sunqa-at-mail.jlu.edu.cn
Date: Mon, 24 Mar 1997 21:18:00 -0800
Subject: Bend contours and Multiply twins

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {33376007.6149-at-mail.jlu.edu.cn}

I'd like to thank Marks, Dorai, Garber, Yi Huang and Ian MacLaren for
their
suggestions to my questions.

Bend contours and multiply twins are possible reasons for the formation
of the pattern
described in my questions. I will find more details on these matters.

Sun Haiping




From: Greg Erdos :      gwe-at-biotech.ufl.edu
Date: Mon, 24 Mar 1997 08:27:53 -0500
Subject: Re: Aclar film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

A commercial source for Aclar is ProPlastics, Inc, P.O. Box 679, Linden, NJ
07036
201-925-5555. They have a minimum order. Several years ago we bought two
pounds of 5mil for $75.00. It will be a lifetime supply.


At 03:55 PM 3/21/97 -0500, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
Scientific Director,
ICBR Electron Microscopy Core Lab
218 Carr Hall Fax: 352-846-0251
University of Florida E-mail: gwe-at-biotech.ufl.edu
Gainesville, FL 32611 http://www.biotech.ufl.edu/~emcl/
Home of the #1 Gators
*****
"Many shall run to and fro, and knowledge shall be increased"
Daniel 12:4





From: Sally Shrom :      sally-at-retina.anatomy.upenn.edu
Date: Mon, 24 Mar 1997 09:52:09 -0500 (EST)
Subject: Re: Aclar film

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Gary,
Aclar is great! Read the paper in Journal of Electron Microscopy Technique
10:77-85 (1988), by Kingsley and Cole. I have used it in epon embedding
protocols. There is
nothing like it. You can get it from Electron Microscopy Sciences. Call
them since it may not be listed in their catalogue. Ask to speak to Stacie
Kirsch.

Aclar is a sturdy thin film, transparent as glass, no detectable
autofluorescence, chemically inert, sections beautifully, and so on and so
on.

Have fun.

Sally

On Fri, 21 Mar 1997, Gary Radice wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Several of you have recommending using Aclar film in your UV polymerization
} protocols. What is Aclar film, and where do you purchase it?
}
} Gary Radice, Associate Professor gradice-at-richmond.edu
} Department of Biology 804-289-8107 (voice)
} University of Richmond VA 23173 804-289-8233 (FAX)
}
}
}





From: johnf-at-geology.wisc.edu
Date: Mon, 24 Mar 1997 10:59:27 -0600
Subject: Electron Microprobe looking for new home

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

(I am posting this for a colleague in France who is not on this list.
Contact him, not me, please.)
==============================

My laboratory has bought recently a SX-100, so I should be happy to find
somebody who might be interested in purchasing our Camebax Electron
Microprobe, which is in good condition.

Here is a short description :
Camebax with 3 WDS and 1 EDS for fully compatible coupled analysis
WDS-EDS (Quantitative analysis on fixed points or lines, X-ray mapping)
WDS : all usual crystals, plus PC1 and PC2 for light elements
EDS : Tracor model, with Be window
Equiped with Back-Scattered Electron detector, and anticontamination
system.
Software : all the software provided by Tracor (Noran) for TN-5500 :
- to control spectrometers, sample holder, and faraday cup
- to perform quantitative analysis (ZAF and PRZ)
- to make standardless analysis (EDS)
- Image analysis (by digitalisation)

The column is from 1976, but everything else is from 1988 (including
spectrometers)

Price = $30 000 USD. (shipping to be paid by purchaser)

For more information, contact:

Michel LAHAYE
ICMCB, Universiti Bordeaux I, Pessac (FRANCE)
Tel 33 5 56 84 62 93
Fax 33 5 56 84 27 61
e-mail : lahaye-at-icmcb.u-bordeaux.fr
==================================================================






From: mboucher-at-pop.isd.net
Date: Mon, 24 Mar 1997 12:13:10 +0000
Subject: Re: LM - point counter

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Jussi:
We could only find Swift point counters when we were in the market
for a new counter for mineralogical applications. They were still in
business 5 years ago when we bought our newer electronic model. There may be other
companies selling similar products. We had an ancient one (} 35
years)that with maintenance the stage worked, but the old electric
pushbuttons were in bad shape. It is dependable and a very adaptable
system. If you get all the gear sets, you can count at any optimum step
size typical of optical microscope use. We used it primarily for
opague ore minerals in polished briquettes. The U.S. office is in San
Jose CA. Fax 408-292-7967. I'm sure there is a closer one to you in
Europe, but I don't have any info on it. You might try looking on the
WWW for a webpage.

I am presently unemployed and have no financial interest in Swift or
anything at present, but would be glad to have some!
================================================
Michael L. Boucher Sr. mboucher-at-isd.net
13345 Foliage Avenue
Apple Valley, MN 55124-5603 Ph 612-432-8836
================================================




From: Ya Chen :      ychen14-at-facstaff.wisc.edu
Date: Mon, 24 Mar 1997 12:55:36 -0600
Subject: Re: preparation/coating of beeswax

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Thu, 20 Mar 1997, Ginger Baker wrote:
}
} Hello all,
}
} I have a PhD student from Brazil who is studying Catolaccus
} grandis. He places a female on top of a beeswax cell which contains
} boll weevil larva. The female deposits an egg on the larva and 2
} weeks later, it matures and burrows its way out of the beeswax
} creating a hole. He would like to study and measure these holes.
}
} However, from past experience I know that beeswax melts when placed
} in my Au/Pd sputter coater. Is there any way to coat the beeswax and
} examine them with an SEM? Or should he try another route?
}
} Thank you in advance,

} Email: lizard-at-okway.okstate.edu
}
Hi Ginger,
I guess the coater you used is a diode type of sputter coater. Heat is
generated during coating. In order to solve your problem, you should to
use a kind of "cooled sputter coater"-- planar magnetron sputter (PMS)
coater. A permanent magnet is posstioned at the center of the cathode to
deflect the electrons away from the specimen.

Ya Chen






From: Ya Chen :      ychen14-at-facstaff.wisc.edu
Date: Mon, 24 Mar 1997 13:01:45 -0600
Subject: Re: preparation/coating of beeswax

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

On Thu, 20 Mar 1997, Ginger Baker wrote:
}
} Hello all,
}
} I have a PhD student from Brazil who is studying Catolaccus
} grandis. He places a female on top of a beeswax cell which contains
} boll weevil larva. The female deposits an egg on the larva and 2
} weeks later, it matures and burrows its way out of the beeswax
} creating a hole. He would like to study and measure these holes.
}
} However, from past experience I know that beeswax melts when placed
} in my Au/Pd sputter coater. Is there any way to coat the beeswax and
} examine them with an SEM? Or should he try another route?
}
} Thank you in advance,

} Email: lizard-at-okway.okstate.edu
}
Hi Ginger,
I guess the coater you used is a diode type of sputter coater. Heat is
generated during coating. In order to solve your problem, you should to
use a kind of "cooled sputter coater"-- planar magnetron sputter (PMS)
coater. A permanent magnet is positioned at the center of the cathode to
deflect the electrons away from the specimen.

Ya Chen


Ya Chen

=========================================================================
\ / Integrated Microscopy Resource (IMR)--
\ / __ an NIH Biomedical Research Resource TEL : 608-263-8481
\/ / / University of Wisconsin-Madison FAX : 608-265-4076
/ / / 1675 Observatory Drive #159 Email1:ychen14-at-facstaff.wisc.edu
/ /__/_ Madison, WI 53706 Email2:chen-at-calshp.cals.wisc.edu
=========================================================================
IMR WWW Home Page: http://www.bocklabs.wisc.edu/imr.html






From: Vickie Frohlich :      vickie-at-vms2.macc.wisc.edu
Date: Mon, 24 Mar 1997 13:40:35 -0600
Subject: Multi-photon Excitation-Symposium and Workshop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

{fontfamily} {param} Courier {/param} {bigger} {bigger} "APPLICATIONS OF
MULTI-PHOTON EXCITATION IMAGING" {/bigger}


A pre-MSA SYMPOSIUM AND SHORT-COURSE


Sponsored by:


Integrated Microscopy Resource

University Wisconsin-Madison


Center for Light Microscope Imaging and Biotechnology=20

Carnegie Mellon University


and=20


Microscopy Society of America


at the=20


SHERATON CITY CENTER, Cleveland, Ohio


Registration Fee: $ 30.00 (Symposium only)

$230.00 (Symposium plus Short-course)


To Register complete form at end of message and submit to: =09

Dawn Volkman

Integrated Microscopy Resource

University of Wisconsin

Madison, WI 53706=20

Phone: (608) 265-3083

Fax: (608) 265-4076

dvolkman-at-students.wisc.edu

www.bocklabs.wisc.edu/imr.html

{color} {param} EEEE,0000,0000 {/param}

{/color} SYMPOSIUM PROGRAM:

SATURDAY 9 AUG 97 - Sheraton Hotel Conference room

8:00- 8:50 Winfried Denk - Lucent Technology/Bell Labs

"Multi-Photon Excitation: From the Beginnings to

Applications in Neuroscience"

8:50- 9:30 Warren Zipfel - Cornell University

"Multi-Photon Excitation of Intrinsic Fluorescence

in Cells and Intact Tissue"

9:30-10:10 David Piston - Vanderbilt University

"Metabolism, Development, and Two-Photon=20

Excitation Microscopy"

10:10-10:30 Coffee Break

10:30-11:10 Stefan Hell - University of Turku, Finland

"Light Microscopy with Sub-100 nm Resolution=20

in Three Dimensions"

11:10-11:50 Enrico Gratton - University of Illinois

"Two-Photon Fluctuation Correlation Spectroscopy"

11:50-12:30 John White - University of Wisconsin

"Multi-Photon Excitation Imaging Applied

to the Study of Developing Embryos"

12:30- 1:30 Lunch

1:30- 2:10 Wayne Knox - Lucent Technology/Bell Labs

"The Path to OEM Femtosecond Sources"

2:10- 2:50 Frank Wise - Cornell University

"Femtosecond-Pulse Sources for Nonlinear

Laser Microscopy"

2:50- 3:10 Coffee Break

3:10- 3:50 Allister Ferguson - University of Glasgow, UK

"User-Friendly Femtosecond Sources for Multi-Photon

Fluorescence Imaging"

3:50- 4:30 Ursula Keller - ETH, Switzerland

"SESAM Devices for Passive Pulse Generation

from 6.5 fs to ns in Solid-State Lasers"

4:30- 5:30 Question/answer session with Saturday speakers

-----------------------------------------------------------------------

SUNDAY 10 AUG 97 - Sheraton Hotel Conference room

8:00- 8:30 Brad Amos - Medical Research Council, UK

"MPEFM: a YLF Laser Applied to Bioimaging"

8:30- 9:00 Hans Gerritsen - Univ. Utrecht, Netherlands

"Fluorescence Lifetime Contrast in Two-photon=20

Excitation Microscopy"

9:00- 9:30 Rafael Yuste - Columbia University

"Two-photon Imaging of Dendritic Spines"

9:30-10:00 Rebecca Williams - Cornell University

"Three-photon Excited Fluorescence Microscopy=20

of Serotonin Release"

10:00-10:15 Coffee Break

10:15-10:45 Steve Potter - California Institute of Technology

"3D Time-lapse Imaging of Hippocampal Slices"

10:45-11:15 Jackie Schiller - Mayo Clinic

"Multi-photon Excitation Uncaging in Rat Brain

Slices"

11:15-11:45 Andy Hargreaves - University of Bristol, UK

"Calcium Imaging with Two-photon Confocal

Microscopy" =20

11:45-12:15 Mark Cannell - University of London, UK

"2PEFM of Calcium in Muscle Cells"

12:15-12:45 Martin Kohler - Karolinska Hospital, Sweden

"2PEFM of Calcium in Pancreatic Beta Cells"

"Islet of Langerhans, Calcium, and Two-photon

Excitation microscopy"

12:45- 1:30 LUNCH

1:30- 4:30 Talks by reps from commercial exhibitors:

Biorad, Nikon, Spectra-Physics, Coherent etc.

=20

4:30- 5:30 Question/answer session with Saturday speakers



{underline} HANDS-ON SHORT COURSE {/underline} for pre-selected
participants will take place

in the Cleveland Convention Center. The short-course will run

concurrent with the afternoon sessions both Saturday and Sunday.


Several multiple-photon systems will be available for "hands-on"

instruction. Each system will vary in its configuration of

scanhead and laser options. Tutorials and discussions will be

provided by IMR staff, commercial exhibitors, and speakers from the

symposium.

Space is limited to 15-20 students. Those requesting admission to

the short-course must apply in writing. A letter outlining their

research interests including a description of their need for

multiple-photon excitation imaging should be submitted to:

Dawn Volkman no later than June 1st.


__________________________________________________________



SYMPOSIUM / SHORT-COURSE REGISTRATION FORM


{bigger} "APPLICATIONS OF MULTI-PHOTON EXCITATION IMAGING"


NAME: SS#:


CITY: STATE: ZIP:

PHONE: FAX:


{/bigger}

REGISTRATION FEE: $ 30.00


Check to MSA enclosed:=20
{/bigger} {/fontfamily} {bigger} {fontfamily} {param} New_York {/param} =03 {/fontfa=
mily} {fontfamily} {param} Courier {/param}


Credit Card Number (Visa or Master Card only): =20


_ _ _ _ - _ _ _ _ - _ _ _ _ - _ _ _ _


Expires: Month _ _ Year _ _


Signature:______________________________________


Print Name on Card:_____________________________


MAIL TO: Dawn Volkman

Integrated Microscopy Resource

University of Wisconsin

Madison, WI 53706=20

Phone: (608) 265-3083

Fax: (608) 265-4076

dvolkman-at-students.wisc.edu

www.bocklabs.wisc.edu/imr.html


SHORT-COURSE REGISTRATION:

Complete the Symposium Registration form with fee and submit with a

letter of application.

Once accepted to the SHORT-COURSE, you will be notified and asked

to submit an additional fee of $200.00 {/fontfamily} {/bigger}






From: ad408-at-detroit.freenet.org (Gilbert T. Groehn)
Date: Mon, 24 Mar 1997 14:37:22 -0500
Subject: LM Need Zeiss & Leitz parts.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I am looking for a source of adapters to convert the
old short barrel Leitz objectives to the newer long barrel
type (170mm tube length).

I also need a set of PLAN APO's for a Zeiss Photomicroscope
in 160mm tube length. Need 4X, 10X, 20X, 40X and 100X O.I.
or anything close in Plan Apo's.

Also need the PHACO 2 and PHACO 3 inserts for a Leitz
Laborlux-11 (type 55 condenser).

Any leads would be most appreciated.

Cordially,
Gil Groehn
(313) 884-1139

--
=============================================================
ULTRAMED, INC. Research Div. ad408-at-detroit.freenet.org
Grosse Pointe Farms, MICH 48236 USA PHONE: 313-884-1139
===============================================================




From: Lesley Weston :      lesley-at-unixg.ubc.ca
Date: Mon, 24 Mar 1997 11:50:10 -0800 (PST)
Subject: Re: Prep of bees wax.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Reprosil is good, and so are any of the other polyvinyl siloxane impression
materials, available from any dental supplier. The resin can be Epon, or
302-1 from Epo-Tek or anything from the EM world.

Lesley Weston


On Fri, 21 Mar 1997, Ronald LHerault wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} An alternative may be replication using Reprosil, a Polyvinyl Siloxane
} from Caulk. It makes a negative which can be transfered to a positive by
} using Spurr low viscosity embedding medium. Detail is there albeit a
} little softened.
}
} Ron
}
}





From: Seth J. Grotelueschen :      sethg-at-CompuServe.COM
Date: Mon, 24 Mar 1997 16:33:12 -0500
Subject: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi,

I was looking at your page and hoped you might help me. I am looking for a
Windows based SEM capture board. We do quite a bit of optical microscope
image capture, but nothing yet on SEMs.

I have heard there are a couple out there, but haven't found them yet!

Thanks
Seth Grotelueschen




From: beebed-at-ere.umontreal.ca (Dwight Beebe)
Date: Mon, 24 Mar 1997 17:35:21 -0400
Subject: Method for Holey grids?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Good afternoon,
I was wondering if those of you who successfully make your own
holey grids for stigmation would care to share the method. I've been using
glcerol and 0.25% Formvar (old bottle) and while I can produce small, nice
looking "holes", on close inspection they're film. There are some true
holes, but few and far between. Any suggestions? Many thanks in advance.


Dwight Beebe E-mail: beebed-at-ere.umontreal.ca
Institut de recherche en biologie vegetale Voice: 514-872-4563
Universite de Montreal FAX: 514-872-9406
4101, rue Sherbrooke est
Montreal, Quebec H1X 2B2
Canada






From: allardlfjr-at-ornl.gov (Larry Allard)
Date: Mon, 24 Mar 1997 23:10:56 -0500
Subject: Re: Method for Holey grids?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dwight:

We historically have made our own holey films, because we could not buy
good holey support films from any manufacturer. Without question, it is an
art, and most of the time a major pain in a posterior region. I would be
happy to send you the detailed instructions we have found to be most
reproducible.

Recently, however, we purchased some holey films from SPI, which have been
uniformly *gorgeous*. They are thin, clean, a large fraction of holes etc.
Probably equivalent to the best I have ever made. I don't recall the
price, but my impression is that I can't make them as cheaply, so why
bother.

Highly recommended.

Larry

PS I have no particular connection to SPI.




} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America






From: Mary Mager :      mager-at-unixg.ubc.ca
Date: Mon, 24 Mar 1997 21:05:20 -0800
Subject: Re: Method for Holey grids?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Dwight,
I know of two methods of preparing holey films. The first is to shake up a
mixture of liquid soap and water and Collodion until it froths, then drop a
drop or two on the surface of distilled water. Lay the grids on top of the
Collodion layer, pick up with a filter paper, dry and carbon coat. Put the
grids in a Jaffe washer with cloroform for 48 hours to remove the Collodion.
The other method, if you have Nucleopore type filters, is to carbon coat a
strip of Nucleopore filter film, place a square of the coated filter on the
grid sitting on a nickel mesh on the surface of the Jaffe washer full of
cloroform, wait 48 hours to dissolve the Nucleopore, dry grids. The holes in
the Nucleopore will be now be holes in the carbon coat.
I must admit I have not tried the first method myself, but someone in my lab
did, years ago.
You wrote:
} I was wondering if those of you who successfully make your own
} holey grids for stigmation would care to share the method. I've been using
} glcerol and 0.25% Formvar (old bottle) and while I can produce small, nice
} looking "holes", on close inspection they're film. There are some true
} holes, but few and far between. Any suggestions? Many thanks in advance.
}
}
} Dwight Beebe E-mail: beebed-at-ere.umontreal.ca
} Institut de recherche en biologie vegetale Voice: 514-872-4563
} Universite de Montreal FAX: 514-872-9406
} 4101, rue Sherbrooke est
} Montreal, Quebec H1X 2B2
} Canada
Good luck,
Mary
Mary Mager
Electron Microscopist
Metals and Materials Eng., UBC
6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel:604-822-5648, fax:604-822-3619
e-mail: mager-at-unixg.ubc.ca





From: Felicity Lawrence :      f.lawrence-at-qut.edu.au
Date: Tue, 25 Mar 1997 14:47:57 +1000 (EST)
Subject: Geranine G sources and information

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

RE: Geranine G

I am looking for any information on the fluorescent dye: Commercial name;
'Geranine G', C.I. name; Direct Red 48,
C.I. number; 14930. I know that it is a monoazo dye with good affinity for
certain proteins, so my question is does anyone have any further information
on Geranine G, such as Excitation and Emission wavelengths, or even
manufacturers or distributors that carry this dye. Failing this would
anyone know of a substitute of similar properties.

My thanks to any and all replies to this request.





From: Xiaoqing Pan :      panx-at-engin.umich.edu
Date: Tue, 25 Mar 1997 00:43:38 -0500
Subject: Positions available at the University of Michigan.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

{fontfamily} {param} Times_New_Roman {/param} {smaller} One Post-Doctoral
Position and one Graduate Assistantships Available at
the {/smaller} {/fontfamily} {smaller} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} University of
Michigan. {/fontfamily} {fontfamily} {param} Times {/param}


{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} We are in
search of a Post-Doctoral Scholar to carry out
research {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} in the areas of
processing and TEM characterization of functional ceramics,
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} including thin
films, thick films and bulk materials. Candidates should have
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} a Ph.D. in a
physics- or materials-related field, with outstanding academic and
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} publication
records. Hands-on experience in the areas of thin thin processing and
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} TEM (HRTEM
and/or AEM) characterization of crystal defects and interfaces in
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} the material is
highly desirable. This position is available immediately for one year,
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} with the
possibility of being renewed for an additional two years. The salary
level {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} will be
commensurate with qualifications and
experience. {/fontfamily} {fontfamily} {param} Times {/param}


{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} We are also in
search of one Graduate Research Associate to pursue Ph.D. in
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} the general
field of functional ceramics, starting Summer/Fall 1997. The
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} candidates
should have a M.S. in a materials-related field, with an excellent
{/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} academic=20
record. The assistantship carries with an attractive stipend
(~16,3k), {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} a full-payment
of tuition, and benefits. {/fontfamily} {fontfamily} {param} Times {/param}


{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} =20
Interested candidates should contact Prof. Pan, preferably by
Email: {/fontfamily} {fontfamily} {param} Times {/param}


{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} Prof. Xiaoqing
Pan {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} Department of
Materials Science and
Engineering {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} University of
Michigan {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} Ann Arbor,
MI-48109-2300 {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} Phone: (313)
647 6822 {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} FAX: (313)
763 4788 {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} Email:=20
panx-at-umich.edu {/fontfamily} {fontfamily} {param} Times {/param}

{/fontfamily} {fontfamily} {param} Times_New_Roman {/param} Web: =20
http://msewww.engin.umich.edu/mse/pan.html {/fontfamily} {fontfamily} {param} Ti=
mes {/param}

{/fontfamily} {/smaller}






From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Tue, 25 Mar 97 00:53:14 -0500
Subject: Bees wax "replicas"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Ron Herault wrote:
-----------------------------------------
An alternative may be replication using Reprosil, a Polyvinyl Siloxane from
Caulk. It makes a negative which can be transfered to a positive by using
Spurr low viscosity embedding medium. Detail is there albeit a little
softened.
-----------------------------------------
Another alternative, possibly an even better choice, would be our own "Wet
Surface Replica Kit", it too is silicone based, but it has been more
optimized for SEM applications. It can be cured into a "negative" much more
quickly, and when converted, the "positive" replicating system is much
easier to use than the "Spurr" approach and on a "per replica basis" is a
lot lower in cost. A demo set of micrographs made from such positive
replicas, on human skin, can be seen in our electronic catalog at the
website mentioned below.

Disclaimer: SPI Supplies manufactures the above mentioned "Wet Replica Kit"
and would have a vested interest in seeing more persons using it.

Chuck

===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: http://www.2spi.com
############################
==================================================




From: Xiaoqing Pan :      panx-at-engin.umich.edu
Date: Tue, 25 Mar 1997 01:24:11 -0500
Subject: Positions available at the University of Michigan.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

One Post-Doctoral Position and one Graduate Assistantships Available at
the

University of Michigan.


We are in search of a Post-Doctoral Scholar to carry out research

in the areas of processing and TEM characterization of functional
ceramics,

including thin films, thick films and bulk materials. Candidates
should have

a Ph.D. in a physics- or materials-related field, with outstanding
academic and

publication records. Hands-on experience in the areas of thin thin
processing and

TEM (HRTEM and/or AEM) characterization of crystal defects and
interfaces in

the material is highly desirable. This position is available
immediately for one year,

with the possibility of being renewed for an additional two years. The
salary level

will be commensurate with qualifications and experience.


We are also in search of one Graduate Research Associate to pursue
Ph.D. in

the general field of functional ceramics, starting Summer/Fall 1997.
The

candidates should have a M.S. in a materials-related field, with an
excellent

academic record. The assistantship carries with an attractive stipend
(~16,3k),

a full-payment of tuition, and benefits.


Interested candidates should contact Prof. Pan, preferably by


Prof. Xiaoqing Pan

Department of Materials Science and Engineering

University of Michigan

Ann Arbor, MI-48109-2300

Phone: (313) 647 6822

FAX: (313) 763 4788

Email: panx-at-umich.edu

Web: http://msewww.engin.umich.edu/mse/pan.html or

http://msewww.engin.umich.edu/mse/pan.html






From: Jim McCarty :      soob-at-teleport.com
Date: Tue, 25 Mar 1997 01:10:03 -0800 (PST)
Subject: Positions available at the University of Michigan.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html






From: Dominique Miller :      dominique-at-ffaltd.demon.co.uk
Date: Tue, 25 Mar 1997 09:04:18 +0000
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

In message {199703241633_MC2-1336-770C-at-compuserve.com} , "Seth J.
Grotelueschen" {sethg-at-CompuServe.COM} writes

Dear Seth

Is is not obvious which country you are from but if you are interested
in someone in the UK try Deben Research who will be able to help you.
If you want the full address and contact no let me know.

They also support our imaging software.

Best regards

Dominique
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

----------------------------------------------------------------------------
|From: Dominique Miller | |
|Marketing Executive | |
|Foster Findlay Associates | Phone: National (0191) 201 2180 |
|Newcastle Technopole | International +44 191 201 2180 |
|Kings Manor | Fax: National (0191) 201 2190 |
|Newcastle upon Tyne | International +44 191 201 2190 |
|NE1 6PA |E-Mail: Dominique-at-ffaltd.demon.co.uk |
|UK | WWW: http://www.demon.co.uk/ffaltd/ |
----------------------------------------------------------------------------




From: veys-at-bota.ucl.ac.be (pascal veys)
Date: Tue, 25 Mar 1997 12:54:37 +0100
Subject: Problem : quantitative morphological analysis ?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear All,

We are studying some ultrastructural features of a new type of epidermal
trichomes on plants.
Trying to define and quantify ulstrastructural changes during the cell
development (from meristematic stage to fully differentiated trichome) we
have to deal with the following questions.
How to start analysing TEM micrographs in order to reach general
conclusions on variations of organelle numbers ?
Are there easy methods to start quantitative morphological analysis ?
Do we have to work with image analysing softwares or not ?
How do you deal with this kind of problem ?
Do you know about books or papers on similar studies that can be used as
references?

Any solutions are welcome...
Thanks !



************************************
Pascal VEYS
Laboratory of Plant Biology
Catholic University of Louvain
Place Croix du Sud 5 (bte 14)
B 1348 Louvain-la-Neuve
Belgium
Phone : 0032 10473004
Fax : 0032 10473471
Email : Veys-at-bota.ucl.ac.be
************************************






From: SGKCCK-at-aol.com
Date: Tue, 25 Mar 1997 07:31:35 -0500 (EST)
Subject: COURSE ANNOUNCEMENT:ULTRAMICROTOMY FOR MATERIALS SCIENCE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

COURSE ANNOUNCEMENT:

ULTRAMICROTOMY ANY CRYO METHODS FOR MATERIALS SCIENCE

PLACE: UNIVERSITY OF PENNSYLVANIA
LABORATORY FOR THE RESEARCH ON THE STRUCTURE OF MATTER
PHILADELPHIA, PA 19104
DATE: JUNE 11-13,1997
SPONSORS: LEICA INC., DIATOME U.S., ELECTRON MICROSCOPY SCIENCES

COURSE SPEAKERS AND INSTRUCTORS: Dr. Tom Malis, Phil Swab, Helmut Gnagi

OVERVIEW:

This workshop will cover the use of Room Temperature and CryoUltramicrotomy
techniques as it applies to Material Science. Included in
the course will be all of the preparation leading up to sectioning, such as,
embedding, trimming and staining on different types of industrial
specimens(polymers, metals, plastics, fibers, rubbers, powders, foils, etc.).
From all of the topics to be covered and in depth review of all of the
different techniques will take place and there will be an emphasis on
"Hands-On" lab time. Participants are encouraged to bring their own samples
as well.
Our goal, upon completion of the course, is that each of the participants
will be able to return to there own lab with a better understanding of the
theory and principles behind ultramicrotomy, and the ability to successfully
perform all of the techniques which we shall cover.

COURSE COST: $1,500.00
The price includes: Hotel accommodations(3 nights), lunch daily, continental
breakfast daily, 1 group dinner, all course supplies and full lab time.

FIRST COURSE DEADLINE:
April 18, 1997

Parties that are interested may either respond by E-Mail to SGKCCK-at-AOL.COM or
Leica's seminar voice mail to 1-800-248-0665 EXT:5010
Contacts are Stacie Kirsch or Ann Korsen.
To confirm your space in the course below is a form that may be filled out
and either E-Mailed to SGKCCK-at-aol.com or faxed to 215-646-8931 ATT: Stacie
Kirsch.
All checks should be mailed to Diatome U.S. P.O. Box 125, Fort Washington, Pa
19034 and made out to Leica, Inc.

WORKSHOP APPLICATION FORM:
Company Name:________________________
Participant's Nmae:______________________
Full Address:____________________________
City, State, Zip:__________________________
Tel #:___________________________________
Fax #:___________________________________
E-Mail:__________________________________
Best Time To Reach:______________________
Area of Interest: Cryo or Room Temperature
Type of Material Working With:______________
Will you be bringing your own sample?___________________What:________

Paying by PO#:________________________(Please mail to P.O. Box)
Check#:_______________________________

If paying by Purchase order number please be sure to include your full
billing address.




From: Jim McCarty :      soob-at-teleport.com
Date: Tue, 25 Mar 1997 05:22:18 -0800 (PST)
Subject: index

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

index





From: john humenansky :      jhumenansky-at-brauncorp.com
Date: Tue, 25 Mar 1997 07:40:25 -0800
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

please subscribe: jhumenansky-at-brauncorp.com




From: Jan Coetzee EM Univ Pretoria :      JANC-at-ccnet.up.ac.za
Date: Tue, 25 Mar 1997 15:55:43 CAT-2
Subject: SEM calibration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear All,

The normal replicas of ruled gratings allow SEM magnification
calibration up to about 50 000x mag. At magnifications higher than
this it is not so easy to find good accurate SEM calibration
specimens.

What SEM calibration specimens are available for use at
50 000 to 1 000 000 x mag in an In-lens FEGSEM?

Regards,



Dr Jan Coetzee
Head: Unit for Electron Microscopy Tel:+27-12-420-2075
University of Pretoria Fax:+27-12-342-1738
Pretoria 0002 Internet:janc-at-ccnet.up.ac.za
South Africa http://www.up.ac.za/academic/electron/emunit1.htm




From: ANDRADY-at-RTI.ORG
Date: Tue, 25 Mar 1997 09:20:43 -0500 (EST)
Subject: SEM calibration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Please unsubscribe




From: Ambrose, Wallace :      wambrose.drc-at-mhs.unc.edu
Date: Tue, 25 Mar 1997 10:15:24 -0500
Subject: nitrogen in titanium eds spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello,
A recent upgrade in EDS instrumentation and detectors has given us the
ability to detect low elements (C, N, O). I have been detecting a peak
that corresponds to nitrogen in all the titanium spectra that I have
acquired. This has occurred using pure titanium (T18) or alloyed titanium
(beta ti). The nitrogen peak ranges from 5-10 weight percent of the
spectra. I know that certain elements can implant in Ti, including C and
N, but these samples have not undergone the heating conditions that would
favor ion implantation. The N peaks do not increase under long e-beam
bombardment. The spectra are collected under routine conditions (15KeV,
recommended WD, 25% deadtime, 100 sec, standardless analysis but the
calibration seems correct). I would appreciate it if anyone could shed
some light on this problem.

Wallace Ambrose
Dental Research Center
Univ. of North Carolina
Chapel Hill, NC





From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Tue, 25 Mar 1997 10:25:17 -0800
Subject: Re: geranine G

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Felicity:

The 9th ed of Conn's Biological Stains offers little on Geranine G, only
a few references. Try the Biological Stain Commission, Dept. of
Pathology, Univ. of Rochester Medical Center, Rochester, NY
(716)-275-2751.

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (908)-235-4583; fax -4029 e-mail: mcauliff-at-umdnj.edu
***************************************************************




From: Naresh Shah :      naresh-at-pop.uky.edu
Date: Tue, 25 Mar 1997 11:29:16 -0500
Subject: Re: nitrogen in titanium eds spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Ambrose, Wallace wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello,
} A recent upgrade in EDS instrumentation and detectors has given us the
} ability to detect low elements (C, N, O). I have been detecting a peak
} that corresponds to nitrogen in all the titanium spectra that I have
} acquired. This has occurred using pure titanium (T18) or alloyed titanium
} (beta ti). The nitrogen peak ranges from 5-10 weight percent of the
} spectra. I know that certain elements can implant in Ti, including C and
} N, but these samples have not undergone the heating conditions that would
} favor ion implantation. The N peaks do not increase under long e-beam
} bombardment. The spectra are collected under routine conditions (15KeV,
} recommended WD, 25% deadtime, 100 sec, standardless analysis but the
} calibration seems correct). I would appreciate it if anyone could shed
} some light on this problem.
}
} Wallace Ambrose
} Dental Research Center
} Univ. of North Carolina
} Chapel Hill, NC

Just a quick thought: are you sure that it is a nitrogen peak and not Ti
L, or Ti L Escape peak?
--

Naresh Shah
Associate Research Professor, University of Kentucky
Department of Chemical and Materials Engineering (CME)
Consortium for Fossil Fuel Liquefaction Science (CFFLS)
533 South Limestone Street, Room 111
Lexington, KY 40508-4005
Phone: (606) 257-5119; FAX: (606) 257-7215; e-mail: naresh-at-pop.uky.edu




From: Alfred Kracher :      akracher-at-iastate.edu
Date: Tue, 25 Mar 1997 10:37:40 -0600
Subject: Ti/N interference

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

In reply to Wallace Ambrose,

} A recent upgrade in EDS instrumentation and detectors has given us the
} ability to detect low elements (C, N, O). I have been detecting a peak
} that corresponds to nitrogen in all the titanium spectra that I have
} acquired.

This is the Ti L-l line (the transition of the M-I to the L-III subshell),
a throroughly nasty interference, even in WDS. Contact me if you have
problems with it.

Alfred

-----------------------------------------
Alfred Kracher
akracher-at-iastate.edu
http://www.public.iastate.edu/~akracher
-----------------------------------------






From: Scott Schwinge :      schwinge-at-fhl.washington.edu
Date: Tue, 25 Mar 97 08:44:13 -0800
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

If your SEM screen accepts a Polaroid camera, Polaroid makes a "MicroCam" and
"PhotoPad" color scanner. Together they retail for about $1000. I have only
seen the ad, and I have no first hand information (Polaroid: 1-800-662-8337).
I have no connection to or investment in Polaroid.

There are other manufacturers of frame-grabbers and videoboards for SEMs, but
alas, I discarded the information.

Scott Schwinge
Friday Harbor Labs
University of Washington


} Hi,
}
} I was looking at your page and hoped you might help me.
} I am looking for a Windows based SEM capture board. We
} do quite a bit of optical microscope image capture, but
} nothing yet on SEMs.
}
} I have heard there are a couple out there, but haven't
} found them yet!
}
} Thanks
} Seth Grotelueschen





From: Michael Knotts :      ph281mk-at-prism.gatech.edu
Date: Tue, 25 Mar 1997 11:46:20 -0500 (EST)
Subject: Looking for veteran SEM service engineers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello Microscopists,

Can anyone suggest names of service engineers with experience on
vintage Cambridge electron microscopes? I have a 1971 Cambridge
Stereoscan S4 SEM in need of service. I'd like to locate an
experienced technician in the metro Atlanta area or the southeastern
US, but I would appreciate all references and information. Thanks,

Michael Knotts
-------------------------------------------------------------
Michael E. Knotts, Ph.D. E-mail: ph281mk-at-prism.gatech.edu
Contributing Editor {The Light Touch} OPTICS & PHOTONICS NEWS
Georgia Tech / School of Physics / Atlanta, GA 30332-0430
Tel: (404) 894-3422 FAX: (404) 894-9958




From: Joseph M. Oparowski, DTN 225-6538, HLO2-3/J09 :      oparowski-at-asdg.enet.dec.com
Date: Tue, 25 Mar 97 11:50:22 EST
Subject: Re: nitrogen in titanium eds spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Although titanium readily attracts nitrogen and oxygen, I think your
problem is related to peak overlap.

A quick look at x-ray tables reveals:

El/line keV

N Ka1&2 0.392

Ti Ll 0.395

Ti La1&2 0.452

O Ka1&2 0.523

The N Ka and Ti Ll peaks cannot be resolved by typical EDS spectrometers,
since you would need a resolution of ~2 ev. To resovle the N Ka and Ti La
peaks you would need a detector resolution of ~50 ev.

Good luck

Joseph




From: Brad Storey :      brad_storey-at-qmgate.fe.anlw.anl.gov
Date: 25 Mar 97 10:46:26 U
Subject: electron-sample interaction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

electron-sample interaction programs
Hello,
I am in need of a Monte Carlo based program (PC or Mac) that models the
interaction of an electron beam with a user defined sample. We have a 200 keV
TEM and an SEM capable of imaging at low voltage; hence, the program should
handle electron transparent samples up to bulk materials and a ~1 to 200 keV
electron beam of variable probe size. I am not only interested in the electron
interaction volume but also in the volume from which many other signals can be
collected, e.g., backscattered electrons, x-rays, secondary electrons, Auger
electrons, etc.

I have seen electron flight simulator for sale ($500), but have little info on
what it provides. What about quality shareware?

Thank you

Brad Storey
Materials Scientist
Argonne National Lab
208-533-7685






From: paulc-at-gps.caltech.edu (Paul K. Carpenter)
Date: Tue, 25 Mar 1997 09:56:25 -0800
Subject: Re: nitrogen in titanium eds spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

X-Sender: paulc-at-mail.gps.caltech.edu
Message-Id: {v02140b00af5dbfc8820c-at-[131.215.67.97]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

Wallace Ambrose wrote,

} A recent upgrade in EDS instrumentation and detectors has given us the
} ability to detect low elements (C, N, O). I have been detecting a peak
} that corresponds to nitrogen in all the titanium spectra that I have
} acquired. This has occurred using pure titanium (T18) or alloyed titanium
} (beta ti). The nitrogen peak ranges from 5-10 weight percent of the

This is not a nitrogen peak. You are seeing the Ti L-series lines, which
exhibit a strong overlap with the Nitrogen K-series lines.

This overlap is bad enough that it cannot really be easily resolved even
using a wavelength-dispersive spectrometer. It is one of the classic
overlaps that makes analysis of Ti alloys for nitrogen problematic.

Nitrogen detection is hampered by strong absorption of Ni Ka by any carbon
that is in the x-ray path, be it in the sample, the carbon coat on the
specimen, oil on the EDS detector window, window material (like diamond,
for example), etc.

These factors make nitrogen measurement a challenge, since one does not
expect to find it in any real concentration unless it is a distinct phase
(like TiN).

Paul


+----------------------------------------------------+
| Paul K. Carpenter paulc-at-gps.caltech.edu |
| Division Analytical Facility |
| Geological and Planetary Sciences MC 100-23 |
| California Institute of Technology |
| Pasadena, CA 91125 |
| 818-395-6126 (X-ray Lab) 818-568-0935 (Dept. FAX) |
+----------------------------------------------------+






From: colijn.1-at-osu.edu (Henk Colijn)
Date: Tue, 25 Mar 1997 13:03:41 -0500
Subject: Re: Method for Holey grids?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dwight,

I use a method by Kuga and Brown from Philips Electron Optics Bulletin
v126, p.19 (1989) to make lacy formvar grids. First a note of caution. I
use dichloroethane as my solvent for the formvar resin. After having much
difficulty in preparing my films, I was told that when exposed to light,
dichloroethane slowly forms HCl in the solution. The HCl destroys the
integrity of the formvar polymer. Solution: store the formvar solution in
a brown bottle in a dark cabinet.

Method:

- I use 0.25% formvar solution in dichloroethane and freshly cleaved mica
sheets as a substrate. The formvar lifts off the mica much easier than
from glass microscope slides.

- Dip the mica into the formvar solution, wick off excess on a paper towel,
then breathe heavily on the mica for about 5 seconds while it is still wet.
The moisture in your breath condenses in the solution. Caution: Don't
inhale!

- When the mica has dried, score the edges and float off on water.

- Place 200 mesh TEM grids on the floating film. The area with the best
holes will appear milky.

- I then take saran wrap, stretch it tightly across the mouth of a small
(~100ml beaker), and press it down at a slight angle onto the floating
formvar film. The film will stick to saran wrap and you can easily pick it
up off the surface of the water.

- After the film has mostly dried, pick up the grids from the saran wrap
"drumhead" and place them on filter paper.

- The film will have many "pseudoholes" that have a thin residual film
across them.

- Following the method of Kuga and Brown, by heating the film to about the
T(g) temperature you can break these holes open. The time and temperature
are critical. I place the bare filter paper in a small lab oven (not in a
petri dish; it has too much thermal mass) at 110C for 12 minutes.

- I then coat both sides with carbon to stabilize the formvar lace.

I can easily make 50-100 grids in an hour (exclusive of the carbon
coating). These grids make wonderful supports for looking at fine
particulate dispersions.

Cheers, Henk

}
} Good afternoon,
} I was wondering if those of you who successfully make your own
} holey grids for stigmation would care to share the method. I've been using
} glcerol and 0.25% Formvar (old bottle) and while I can produce small, nice
} looking "holes", on close inspection they're film. There are some true
} holes, but few and far between. Any suggestions? Many thanks in advance.
}
}
} Dwight Beebe E-mail: beebed-at-ere.umontreal.ca
} Institut de recherche en biologie vegetale Voice: 514-872-4563
} Universite de Montreal FAX: 514-872-9406
} 4101, rue Sherbrooke est
} Montreal, Quebec H1X 2B2
} Canada

Henk Colijn colijn.1-at-osu.edu OSU Campus Electron Optics Facility
-------------------------------------------------------------------
Prosperity is the blessing of the Old Testament; Adversity is the blessing
of the New. Francis Bacon, "Of Adversity."






From: Sanford Simon :      simon-at-rockvax.rockefeller.edu
Date: Tue, 25 Mar 1997 13:56:12 -0500
Subject: teflon o-rings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Does anyone know of a source for teflon o-rings with a thickness of {100 uM?

Thanks,
Sandy Simon

Sanford M. Simon
Laboratory of Cellular Biophysics
Box 304
Rockefeller University
1230 York Avenue
New York, N.Y. 10021
212-327-8130 (voice)
212-327-8022 (fax)
simon-at-rockvax.rockefeller.edu (e-mail)







From: Donna Wagahoff :      DWAGAHOF-at-wpsmtp.siumed.edu
Date: Tue, 25 Mar 1997 13:06:33 -0600
Subject: Current phone # for Energy Beam Sciences

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Does anyone have a current phone number for Energy Beam Sciences
or whichever company has taken over Polaron? I have a question about
our Polaron film thickness monitor.
Thanks for your help.

Donna Wagahoff
217-782-0898
fax 217-524-3227




From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Tue, 25 Mar 1997 19:26:45 +0000
Subject: Re: SEM calibration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Dear All,
}
} The normal replicas of ruled gratings allow SEM magnification
} calibration up to about 50 000x mag. At magnifications higher than
} this it is not so easy to find good accurate SEM calibration
} specimens.
}
} What SEM calibration specimens are available for use at
} 50 000 to 1 000 000 x mag in an In-lens FEGSEM?
}
} Regards,
}
}
}
} Dr Jan Coetzee

In biological TEM, colloidal gold in a variety of immunological labeling
techniques. This gold has fairly well defined size distributions - the
smallest available has a mean size of 1nm. If you check the EM supplies
catalogues (Agar, SPI, Ted Pella, etc), you should find it - personally,
I've never tried it, but I don't see any reason why it couldn't be used
(although you might need to remove some antibodies).

The other approach is to use the same techniques that is used in TEM, where
ther is a bit of a gap between gratings and lattice resolution specimens.
Basically, you work up step by step, going from a lower mag, where the
grating gives you a calibration, identify a number of obvious features,
then go up in mag and relocate the same features - they got to be the same
distance apart. A bit tedious and not statistically very accurate, +/- 5%
at best, but its better than nothing.

Regards,
Larry Stoter






From: Henrik Kaker :      Henrik.Kaker-at-guest.arnes.si
Date: Tue, 25 Mar 1997 22:06:07 +0100
Subject: Re: electron-sample interaction

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Brad Storey wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} electron-sample interaction programs
} Hello,
} I am in need of a Monte Carlo based program (PC or Mac) that models the
} interaction of an electron beam with a user defined sample. We have a 200 keV
} TEM and an SEM capable of imaging at low voltage; hence, the program should
} handle electron transparent samples up to bulk materials and a ~1 to 200 keV
} electron beam of variable probe size. I am not only interested in the electron
} interaction volume but also in the volume from which many other signals can be
} collected, e.g., backscattered electrons, x-rays, secondary electrons, Auger
} electrons, etc.
}
} I have seen electron flight simulator for sale ($500), but have little info on
} what it provides. What about quality shareware?
}
} Thank you
}
} Brad Storey
} Materials Scientist
} Argonne National Lab
} 208-533-7685

On my laboratory web site are informations about almost all public
domain and commercial Monte Carlo programs for PC and Mac's.

http://www2.arnes.si/guest/sgszmera1/monte.html

--
Henrik Kaker
SEM-EDS Laboratory, Metal Ravne d.o.o., Slovenia,
Tel: +386 602 21 131 Fax: +386 602 20 436
mailto:Henrik.Kaker-at-guest.arnes.si
http://www2.arnes.si/guest/sgszmera1/index.html
Microscopy Vendors Database:
http://www2.arnes.si/guest/sgszmera1/vendors.html
http://www.kaker.com/mvd/vendors.html




From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Tue, 25 Mar 1997 19:12:55 +0000
Subject: Re: nitrogen in titanium eds spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Hello,
} A recent upgrade in EDS instrumentation and detectors has given us the
} ability to detect low elements (C, N, O). I have been detecting a peak
} that corresponds to nitrogen in all the titanium spectra that I have

snips

} calibration seems correct). I would appreciate it if anyone could shed
} some light on this problem.
}
} Wallace Ambrose
} Dental Research Center
} Univ. of North Carolina
} Chapel Hill, NC

Titanium L edges at around 0.45 keV?

Reagrds,
Larry Stoter







From: Glenn Poirier :      glennp-at-eps.mcgill.ca
Date: Tue, 25 Mar 1997 15:27:56 -0500
Subject: Electron Channeling contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Good afternoon everyone.

Can anyone refer me to a basic reference on electron channeling =
contrast, or failing that, give me a few hints on optimal conditions for =
observing this type of contrast ?=20

TIA=20

Glenn

Glenn Poirier Phone (514) 398 6774
Electron Microprobe Laboratory Fax (514) 398 4680
Earth and planetary Science=20
Mcgill University

THERE ARE THREE SIDES TO EVERY STORY:
yOURS, MINE AND THE TRUTH





From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Wed, 26 Mar 1997 09:52:46 GMT+1200
Subject: Peak Overlaps

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Talking of peak overlaps, is anyone out there capable of reliable
accurate analysis of V in the presence of large amounts of Ti, using:

a WDS

b EDS?

I ask because I concluded recently that I couldn't, with my EDS-only
system, analyse for V in titanomagnetites, and that it probably
wasn't possible even with WDS.
If I ask my system to look for V in my standard Rutile (TiO2), it
always finds a few tenths of a %, even though my software
(LINK ZAF4/FLS) strips out Ti Kb along with the Ka. I have no way of
telling whether this V is genuine or not, but I suspect that it
isn't, as my supposedly-pure Ti metal standard also gets credited
with a small amount of V, which I doubt.
I was recently given another rutile standard, and, blow me down, it
is reputed to have 0.4% V by WDS analysis.
Anyone got any feeling for whether this is likely to be true?

cheers

Ritchie

Ritchie Sims phone: 64 9 3737599 ext 7713
Department of Geology fax: 64 9 3737435
University of Auckland
Private Bag 92019
Auckland
New Zealand




From: Ambrose, Wallace[SMTP:wambrose.drc-at-mhs.unc.edu]
Date: Tue, 25 Mar 1997 17:06:17 -0500
Subject: nitrogen in titanium eds spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Wallace:
This is most likely the Ti Ll line overlapping N Ka. Beware of overlaps in the light element energy region!

Jim
*.*.*.*.*.*.*.*.*.*.*.*.*.*.*.*.*.*.*.*.*
James J. McGee (jmcgee-at-sc.edu)
Dept. of Geological Sciences
University of South Carolina (803) 777-6300 (Office)
Columbia, SC 29208 (803) 777-6610 (Fax)


----------

Hello,
A recent upgrade in EDS instrumentation and detectors has given us the
ability to detect low elements (C, N, O). I have been detecting a peak
that corresponds to nitrogen in all the titanium spectra that I have
acquired. This has occurred using pure titanium (T18) or alloyed titanium
(beta ti). The nitrogen peak ranges from 5-10 weight percent of the
spectra. I know that certain elements can implant in Ti, including C and
N, but these samples have not undergone the heating conditions that would
favor ion implantation. The N peaks do not increase under long e-beam
bombardment. The spectra are collected under routine conditions (15KeV,
recommended WD, 25% deadtime, 100 sec, standardless analysis but the
calibration seems correct). I would appreciate it if anyone could shed
some light on this problem.

Wallace Ambrose
Dental Research Center
Univ. of North Carolina
Chapel Hill, NC







From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Tue, 25 Mar 1997 17:15:56 -0400
Subject: RE:Holey Carbon films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

The question of how to make holey carbon films came up on the list server
two years ago. At that time John Gabrovsek (gabrovj-at-ccsmtp.ccf.org) gave
the following reference: Baumeister & Seredynsky, Micron 1976, Vol 7, p.
49, and Jane Fagerland (fagerland.jane-at-igate.abbott.com)gave this one:
Elsner, Proceedings, 29th EMSA Meeting, p. 460. Both methods were said to
work satisfactorily. You might try contacting these people for more
details.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: A Wilson :      awilson-at-aw.u-net.com
Date: Tue, 25 Mar 1997 22:25:15 +0000
Subject: RE:Holey Carbon films

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

subscribe





From: GANTZ-at-med-biophd.bu.edu
Date: Tue, 25 Mar 1997 13:08:08 -0400 (EDT)
Subject: Quantitative morphological analysis

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear pascal veys:
Here are some basic references that may help you get started:
Weibel, E.R. 1979. Stereological methods, vol 1. Practical Methods for
biological morphometry. Academic Press, New York.

Weibel, E.R., "Stereological Techniques for Electron Microscopic
Morphometry". In Principles and Techniques of Electron Microscopy,
Vol. 3, M.A. Hayat, ed. 1973. pp 239-296.

Loud, A.V., 1968. A quantitative stereological description of the
ultrastructure of normal rat liver parenchymal cells. Journal of Cell
Biology, Vol 37:27-46.

Don Gantz
Boston Univ Med School
gantz-at-med-biophd,bu.edu




From: Pete Lander :      pete-at-fenland.demon.co.uk
Date: Tue, 25 Mar 1997 22:28:19 +0000
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Seth,
JEOL in Scandinavia market a system called Semafore, their Web page may
be worth a look as it has sample software to download. email me if you
would like more information.

all the best
Pete Lander (JEOL UK)




} -----------------------.
} }
} } Hi,
} }
} } I was looking at your page and hoped you might help me. I am looking for a
} } Windows based SEM capture board. We do quite a bit of optical microscope
} } image capture, but nothing yet on SEMs.
} }
} } I have heard there are a couple out there, but haven't found them yet!
} }
} } Thanks
} } Seth Grotelueschen
}

Return messages:
Home Work
email pal-at-fenland.demon.co.uk pal-at-jeolsys.demon.co.uk
phone 01354 661413 01707 377117
fax phone first 01707 373255




From: WARRENJ1-at-cliffy.polaroid.com
Date: 3.25.97 11:44 AM
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The MicroCam slides onto the eyepiece of a optical microscope. It doesn't work
with a SEM camera port.
John D. Warren
Area Sales Manager
Digital Products
Polaroid Corporation

4525 Leonard Parkway
Richmond, Virginia 23221-1809

804 254 1011
804 254 1013 Fax
warrenj1-at-polaroid.com



______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

If your SEM screen accepts a Polaroid camera, Polaroid makes a "MicroCam" and
"PhotoPad" color scanner. Together they retail for about $1000. I have only
seen the ad, and I have no first hand information (Polaroid: 1-800-662-8337).
I have no connection to or investment in Polaroid.

There are other manufacturers of frame-grabbers and videoboards for SEMs, but
alas, I discarded the information.

Scott Schwinge
Friday Harbor Labs
University of Washington


} Hi,
}
} I was looking at your page and hoped you might help me.
} I am looking for a Windows based SEM capture board. We
} do quite a bit of optical microscope image capture, but
} nothing yet on SEMs.
}
} I have heard there are a couple out there, but haven't
} found them yet!
}
} Thanks
} Seth Grotelueschen




From: Bob_Citron-at-cc.chiron.com (Bob Citron)
Date: 3/25/97 10:15 AM
Subject: nitrogen in titanium eds spectrum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Mime-Version: 1.0

Hi Wallace;

Don't feel alone. Last year, I tried to do a quantitative analysis of 6/4
titanium/vanadium by EDS. The responses you have received so far explain
the problem - peak overlaps. I had the additional factor of a V L-alpha
emission at 0.51 keV. I opted for emission spectroscopy at an outside lab.

Regards,

-Bob
****************************
Bob Citron
Chiron Vision
Claremont, CA
Bob_Citron-at-cc.chiron.com
****************************

______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hello,
A recent upgrade in EDS instrumentation and detectors has given us the
ability to detect low elements (C, N, O). I have been detecting a peak
that corresponds to nitrogen in all the titanium spectra that I have
acquired. This has occurred using pure titanium (T18) or alloyed titanium
(beta ti). The nitrogen peak ranges from 5-10 weight percent of the
spectra. I know that certain elements can implant in Ti, including C and
N, but these samples have not undergone the heating conditions that would
favor ion implantation. The N peaks do not increase under long e-beam
bombardment. The spectra are collected under routine conditions (15KeV,
recommended WD, 25% deadtime, 100 sec, standardless analysis but the
calibration seems correct). I would appreciate it if anyone could shed
some light on this problem.

Wallace Ambrose
Dental Research Center
Univ. of North Carolina
Chapel Hill, NC





From: Donna Wagahoff :      DWAGAHOF-at-wpsmtp.siumed.edu
Date: Tue, 25 Mar 97 19:11:18 CST
Subject: Current phone # for Energy Beam Sciences

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Does anyone have a current phone number for Energy Beam Sciences
or whichever company has taken over Polaron? I have a question about
our Polaron film thickness monitor.
Thanks for your help.

Donna Wagahoff
217-782-0898
fax 217-524-3227




From: =?ISO-8859-1?Q?Miguel_Angel_S=E1nchez?= :      misanche-at-infosel.net.mx
Date: Tue, 25 Mar 97 19:20:00 PST
Subject: Measuring textiles with microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi there:
My name is Miguel Sanchez
my problem is the following
As part of my business I receive
micron rated textiles, that is nylon mesh
that has a determined distance between
threads.
Sometimes is very difficult to determine
that the textile is inside tolerances or
not.
I use an optical microscope to do measure them
my question is:
What do I need to measure the textiles
with the use of a camera and a computer?
I want to use a camera so we can agree
on what we are measuring.
The solution has to be cost effective. Very cost
effective.
Your comments will be highly appreciated
Thanks






From: RCHIOVETTI-at-aol.com
Date: Tue, 25 Mar 1997 20:22:16 -0500 (EST)
Subject: Re: Current phone # for Energy Beam Sciences

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi Donna,

You can reach Energy Beam Sciences at the following number:

} } 1-800-992-9037 { {

You can also visit their web page at the URL: http://www.ebsciences.com/

Best of luck to you.

Bob Chiovetti




From: Doug Johnson :      djohnson-at-skfcm.com
Date: Tue, 25 Mar 1997 17:25:59 -0800
Subject: subscription

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Subscribe Microscopy Djohnson-at-skfcm.com




From: Ji Yan Dai :      jyd571-at-nwu.edu
Date: Tue, 25 Mar 1997 20:29:49 -0600
Subject: modulated structure in ZrO2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

while studing ZrO2 thin film on sapphire substrate, I found a kind of
modulated structure like a wave near the interface region (It's not a
morie fringe). This kind of structure is very similia to the
pre-martensitic sructure found in Ni-Al system. The lattice spacing in
the modulated region is the same as (111) d-spacing, while in the uper
region d=d(002). The film has a (001)//(0001) prefer orientation with
sapphire. What I want to know is:
1. How the modulated structure formed, and how (111)//(0001) orientation
can change to (001)//(0001)? Can the modulated region comes from a
martensitic transformation?
2. Is there any report about modulated structure happend in ZrO2?




From: Fred Pearson :      eoptics-at-mcmail.cis.mcmaster.ca
Date: Wed, 26 Mar 1997 11:54:52 -0500 (EST)
Subject: EM 300 Parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





We have a Specimen Injection Rod and 4, PW 6145/00 3 mm. sample holders
for an Philips EM 300 Biological Microscope available.
Are far as I know, these have never been used because there is no evidence
of contamination or wear on the rod or holders.

If you are interested, make an offer.


Fred Pearson
McMaster University
Brockhouse Institute for Materials Research
Hamilton Ontario
Canada L8S 4M1

Phone: (905) 525-9140 ext. 24609
Fax: (905) 521-2773







From: Fred Pearson :      eoptics-at-mcmail.cis.mcmaster.ca
Date: Wed, 26 Mar 1997 17:02:00 -0500 (EST)
Subject: EM 300 Parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a Specimen Injection Rod and 4, PW 6145/00 3 mm. sample holders tips
for a Philips EM300 Biological Microscope for sale.

Are far as I know these have never been used because there is no evidence
of contamination or wear on the rod or holder.

If you are interested, make an offer.

email: eoptics-at-mcmail.CIS.McMaster.CA


********************************************************
Fred Pearson
Brockhouse Institute for Materials Research
McMaster University
1280 Main St. West
Hamilton, Ontario
Canada L8S 4M1

********************************************************




From: Bill Hardy :      bhardy-at-qtmsys.com
Date: Wed, 26 Mar 1997 09:02:06 -0500
Subject: Wanted Old Journals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Wanted: Back Copies of
Microbeam Analysis - 19XX
SEM - 19XX

I am trying to build up a reference library and would be interested in
acquiring back copies of journals with relevance to EDS/WDS spectroscopy.

Please contact Bill Hardy (bhardy-at-qtmsys.com) if you have back issues
which you would like to get rid of or sell.

Thanks all,


*************************************************
* Bill Hardy, President *
* American Nuclear Systems, Inc. *
* 12633 Red Canyon Road *
* Knoxville, TN 37922 *
* (423) 671-0292 FAX: (423) 671-0293 *
* WWW.qtmsys.com Email: bhardy-at-qtmsys.com *
*************************************************





From: tongmuan :      tongmuan-at-loxinfo.co.th
Date: Thu, 27 Mar 1997 12:04:17 +0700
Subject: Help with maintenance!!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I own an old Leitz Wetzlar bi-microscope, and have no manual for
maintenance, instructions, etc.
Anyone here able to provide some pointers or reference?
Many thanks.
Frank G Anderson
745 Sipsiri Soi 3
Myang, Korat 30000
Thailand




From: Dave Emmitt :      dremmitt-at-ameritech.net
Date: Wed, 26 Mar 1997 12:40:37 -0500
Subject: Re: Measuring textiles with microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Miguel Angel S=E1nchez wrote:
} =20
} -----------------------------------------------------------------------=
-
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Co=
m
} -----------------------------------------------------------------------.
} =20
} Hi there:
} My name is Miguel Sanchez
} my problem is the following
} As part of my business I receive
} micron rated textiles, that is nylon mesh
} that has a determined distance between
} threads.
} Sometimes is very difficult to determine
} that the textile is inside tolerances or
} not.
} I use an optical microscope to do measure them
} my question is:
} What do I need to measure the textiles
} with the use of a camera and a computer?
} I want to use a camera so we can agree
} on what we are measuring.
} The solution has to be cost effective. Very cost
} effective.
} Your comments will be highly appreciated
} Thanks
You need a reticule and a stage micrometer. The reticule inserts into an
eyepiece and provides a grid. The stage micrometer allows for
measurement of each increment.=20

You can obtain from Klarmenn Ruling, Inc. Manchester NH (603) 424-2401.

You'll need to know the ID of the eyepiece and what measurement scale
you want to use.




From: leibest-at-acpub.duke.edu (Leslie Eibest)
Date: Wed, 26 Mar 1997 08:27:29 -0500
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Seth;
Ted Pella offers Printerface for Windows, an image capture system
that generates digital images from analog SEM images (I think the board is
manufactured by Gatan). I haven't tried it yet, but I've heard it works
well.
(I have no financial interest in Pella or Gatan). Pella's phone # is (800)
637-3526 (USA).

Leslie Eibest
Zoology Dept., Box 90325
Duke University
(919) 684-2547
leibest-at-duke.edu






From: Dave Emmitt :      dremmitt-at-ameritech.net
Date: Wed, 26 Mar 1997 12:31:36 -0500
Subject: Re: teflon o-rings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Sanford Simon wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Does anyone know of a source for teflon o-rings with a thickness of {100 uM?
}
} Thanks,
} Sandy Simon
}
} Sanford M. Simon
} Laboratory of Cellular Biophysics
} Box 304
} Rockefeller University
} 1230 York Avenue
} New York, N.Y. 10021
} 212-327-8130 (voice)
} 212-327-8022 (fax)
} simon-at-rockvax.rockefeller.edu (e-mail)
You can try Roger Zatkoff Company -at- (810) 478-2400, Farmington Hills,
MI. They carry nothing but O Rings.




From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Wed, 26 Mar 1997 16:52:58 -0400
Subject: :Addr for Bob Baier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Bob Baier at the Univ of Buffalo apparently saw my name on this list, and
sent me a personal message about a week ago. I have since tried several
times to reply to him using the address: baier-at-ubvms.cc.buffalo.edu -
(which I copied directly from the original message I received from him),
but keep getting rejections due to 'local delivery' problems. Does anyone
happen to know a better address for him, or have suggestions on how to get
through to him with this one?

TIA

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Wed, 26 Mar 1997 16:15:54 -0400
Subject: RE:EChanneling Contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Chapter 3 of the book "Advanced Scanning Electron Microscopy & X-ray Micro
Analysis" by Newberry, Joy, et. al, Plenum Press 1986 is entitled "Electron
Channeling Contrast in the SEM", and ought to provide the information you
are looking for.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: GIDDINGS THOMAS H :      giddings-at-spot.colorado.edu
Date: Wed, 26 Mar 1997 13:03:25 -0700 (MST)
Subject: freeze-fracture/double replica

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'd be pleased to hear from people who are working with the BAL-TEC
BAF-060 freeze-etch system, if nothing else, to swap stories, experiences
and ideas. I'm thinking about modifying or redesigning the double replica
specimen table, because the BAL-TEC (Technotrade) model looks too
difficult to load without pre-fracturing the sample. Has anyone had good
luck with the off-the-shelf model? or come up with a better design? The
holders with 3 individual sample slots used on older Balzers systems
seemed to work well, but to my knowledge that design is not available for
the BAF 060.

Thanks for any feedback,
Tom

Thomas H. Giddings Tel: (303) 492-8402
MCDB Electron Microscopy Service Fax: (303) 492-7744
University of Colorado giddings-at-spot.colorado.edu
Boulder, CO 80309-0347






From: Mary Mager :      mager-at-unixg.ubc.ca
Date: Wed, 26 Mar 1997 20:06:11 -0800
Subject: Re: Electron Channeling contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Glenn,
From what I can remember, the few times I have see channeling contrast on
the SEM, it is what shows up when other sources of contrast are missing. The
sample must be electropolished (to eliminate surface deformation), smooth
(to eliminate SE contrast), homogeneous (to eliminate BS contrast), then
crank the contrast way up. I think the BS detector works best. High voltage
helps. Albert Curzon of the Physics Department of Simon Fraser University in
Burnaby, B.C. Canada has done SEM studies of magnetic domain, which is
imaged in a similar way.
You wrote:

}
} Can anyone refer me to a basic reference on electron channeling contrast,
or failing that, give me a few hints on optimal conditions for observing
this type of contrast ?
}
} TIA
}
} Glenn
}
} Glenn Poirier Phone (514) 398 6774
} Electron Microprobe Laboratory Fax (514) 398 4680
} Earth and planetary Science
} Mcgill University
Regards,
Mary
Mary Mager
Electron Microscopist
Metals and Materials Eng., UBC
6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel:604-822-5648, fax:604-822-3619
e-mail: mager-at-unixg.ubc.ca





From: Nestor J. Zaluzec :      zaluzec-at-Sparc5.Microscopy.Com
Date: Wed, 26 Mar 1997 23:23:07 -0500
Subject: Microscopy is Back ON-LINE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Colleagues,

I picked up the following message which was posted to another newsgroup
(sci.techniques.microscopy). Anyone who can help, please contact J.L.
Beauchamp directly at Caltech.

Thanks.

Bob Chiovetti

} } } } } } } } } } } } } } } } } } } } } } } } } { { { { { { { { { { { { { { { { { { { { { { { {


G'day All...

If you haven't noticed it the Microscopy Listserver was
off-line for most of the day today.

This note is just to head off a host of messages asking why
people were not able to send/receive Email and/or login to various
MSA WWW sites.

Network Operations Center and Ameritek Communications had
a major fault in the lines to this site. We finally came back
up on-line ~ 11 PM CST . A technical genius at the main switching
center basically didn't know what he was doing and incorrectly
reconfigured a central router about 11 AM this morning.
It took lots of phone calls to get it sorted out.

All the local systems here were running, they just could not talk
to anyone on the outside world. Well not quite true, I still managed
to reconfigure my backup modem connections and send a couple
of messages...

Cheers,

Nestor
Your Friendly Neighborhood SysOp






From: Marc Friedman :      marc-at-accumed.com
Date: Wed, 26 Mar 1997 14:08:59 -0600
Subject: Open Position for Technical Manager

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

AccuMed International, Inc., has an immediate opening for a technical =
manager to oversee scientific and technical projects involved in the =
development, scientific and clinical testing, and production of =
computerized medical laboratory instruments and software-based =
applications. The successful candidate will carry out day-to-day =
activities involved in shepherding new instruments from R&D to finished =
product. The position requires a minimum of a Bachelor's degree and a =
strong scientific/technical background, with at least 3-5 years =
experience in developing, marketing or supporting imaging workstations =
and software-based applications for light microscopy.
Specific areas of responsibility will include:
- managing the development of new hardware and software products
- establishing and maintaining timelines for delivery of prototype =
instruments and finished products, including software-based applications
- coordinating activities in various departments including R&D, =
engineering, software development and manufacturing
- assisting with applications to regulatory agencies

AccuMed International, headquartered in Chicago, is a global laboratory =
diagnostic products company and an emerging leader in the field of =
automated cytopathology. EOE.

Please respond via mail, FAX or email (NOT telephone) to:

Marc M. Friedman, Ph.D.
Director, Scientifc and Technical Affairs
AccuMed International, Inc.
900 N. Franklin, Suite 401
Chicago, IL 60610
Tel: (312) 642-9200
FAX: (312) 642-8684
email: marc.friedman-at-accumed.com





From: Woody.N.White-at-mcdermott.com
Date: 3/25/97 2:27 PM
Subject: Electron Channeling contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

It is not uncommon for me to employ electron channelling contrast.
Almost all of my experience is with stainless steels, nickel
superalloys (Inconels), Zirconium alloys, and uranium compounds. The
ecc is a weak signal and is not always easy to produce.

A few ideas....

Use BSE imaging with large beam currents and high BSE signal gain
(contrast). It may be necessary let any response from areas of higher
or lower Z than the matrix of interest go saturated black and/or white
to achieve this.

Use a "normal" incident beam (0 degrees tilt)

Contrary to some advice I have recieved, I find that lower beam
voltages (10 kv) work better than higher (20-30 kV). I have no proof,
but suspect that the lower penetration depth images the surface grains
without "confusing and diluting" the image with BSE returns from
sub-surface grains with different orientations.

Surface preparation is VERY important. A very well polished surface, free
from
surface damage is required or the signal will be obscured. Some materials
are
easier than others to prepare. On occation, I have had to send samples back
to
our met-lab several times before a satisfactory surface is available. A
very
light "attack" polish (not really an etch) can be helpful during final
polish
for some materials. I have wanted to try a light ion beam cleaning, but
don't
have one.

Incident beam angle (with specimen at 90 nominal) changes resulting from the

raster can strongly affect contrast. Experiment with working distance vs.
magnification to see what works best for you. Along this line... Don't
expect
to be able to generate multiple image mosaics which match well. A grain
that is
dark on one edge of an image may will be light when the stage is translated
one
frame over because the incident angle is nolonger the same.


Hope this helps...

Woody White

http://www.geocities.com/capecanaveral/3722

BTW something happened to my email pgm half way through??? Hope this
dosen't
"run over" line length too much!

______________________________ Reply Separator
_________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Good afternoon everyone.

Can anyone refer me to a basic reference on electron channeling
contrast,
or failing that, give me a few hints on optimal conditions for observing
this type of contrast ?

TIA

Glenn

Glenn Poirier Phone (514) 398 6774
Electron Microprobe Laboratory Fax (514) 398 4680
Earth and planetary Science
Mcgill University

THERE ARE THREE SIDES TO EVERY STORY:
yOURS, MINE AND THE TRUTH




From: Sara Miller :      saram-at-acpub.duke.edu
Date: Wed, 26 Mar 1997 10:06:38 -0500 (EST)
Subject: Re: Looking for veteran SEM service engineers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Tue, 25 Mar 1997, Michael Knotts wrote:

} Date: Tue, 25 Mar 1997 11:46:20 -0500 (EST)
} From: Michael Knotts {ph281mk-at-prism.gatech.edu}
} To: microscopy-at-Sparc5.Microscopy.Com
} Subject: Looking for veteran SEM service engineers
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.
}
} Hello Microscopists,
}
} Can anyone suggest names of service engineers with experience on
} vintage Cambridge electron microscopes? I have a 1971 Cambridge
} Stereoscan S4 SEM in need of service. I'd like to locate an
} experienced technician in the metro Atlanta area or the southeastern
} US, but I would appreciate all references and information. Thanks,
}
} Michael Knotts
} -------------------------------------------------------------
} Michael E. Knotts, Ph.D. E-mail: ph281mk-at-prism.gatech.edu
} Contributing Editor {The Light Touch} OPTICS & PHOTONICS NEWS
} Georgia Tech / School of Physics / Atlanta, GA 30332-0430
} Tel: (404) 894-3422 FAX: (404) 894-9958
}

Yes, Hugh Whitaker (WHIT) used to keep our ancient Cambridge Stereoscan
going with corks and paper clips from the local electronics store when he
couldn't get parts from Cambridge. I think he is retired now, but he is
a wealth of information, and being retired, may be available to help you
out. He used to live in Raleigh, but I think he may have moved to the
coast of NC. His daughter, Sharon Drew, works at Medical Univ. Hosp. in
Charleston 803 792 4157. Perhaps she can get in touch with him for you.
Sara

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735





From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Wed, 26 Mar 1997 08:39:24 +0000
Subject: Re: Peak Overlaps

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} Talking of peak overlaps, is anyone out there capable of reliable
} accurate analysis of V in the presence of large amounts of Ti, using:
}
} a WDS
}
} b EDS?
}
} I ask because I concluded recently that I couldn't, with my EDS-only
} system, analyse for V in titanomagnetites, and that it probably
} wasn't possible even with WDS.
} Ritchie
}
} Ritchie Sims phone: 64 9 3737599 ext 7713
} Department of Geology fax: 64 9 3737435
} University of Auckland
} Private Bag 92019
} Auckland
} New Zealand

EELS on a TEM, with energy resolution around 1-2 eV, pretty easily
separates Ti and V, using the L edges, and there is little problem of
confusion with N K edge. Accuracy of quantitation won't be very good but it
might allow you to decide if you really do have a small amount of V in Ti.

However, what is known as 'Sod's Law' ensures that the O-K edge sits almost
exactly at the energy of the V-L edge! If you have strongish V (or O)
edges, then you can distinguish O-K from V-L on the basis of edge shape,
but if you've got that level of V, you can probably separate V and Ti by
EDX anyway!!

Regards,
Larry Stoter






From: ScottE57-at-aol.com
Date: Wed, 26 Mar 1997 08:37:48 -0500 (EST)
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Seth, for your optical microscope image capture, we have a program called
"Image Central" that will archive and database all of your digital images and
let your acquire them directly from a variety of sources. Please contact me
fro further info and I will send a package of literature out to you.

Scott E. Berman
Advanced Imaging Concepts, Inc.
Princeton, NJ
Phone: (609) 921-3629 x26
Fax: (609) 924-3010
e-mail : Scott E57-at-aol.com




From: Chris Johns :      c.johns-at-student.qut.edu.au
Date: Thu, 27 Mar 1997 17:51:24 +1000 (EST)
Subject: Nitrogen Drying

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have just read an artical that includes in its protocol for fluorescent
dyeing, blot drying of the samples, then drying under a stream of nitrogen.

My question was what is the benefit of drying under a stream of nitrogen as
opposed to air drying? My first thought was perhaps to reduce any oxidation.
Can anyone shed any light on this matter.

Thanks,

Chris.
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Chris Johns
Queensland University of Technology
2 George St.
Brisbane QLD 4001.
Email: {c.johns-at-student.qut.edu.au}
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: Paul Vanderlinden :      orion-at-infoboard.be
Date: Thu, 27 Mar 1997 11:06:30 +0100 (MET)
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please find hereafter the major characteristics of our grabbing system for
SEMs: Orion 4.2 for Windows.

=B7 Digitizes in any size between 16x16 and 4096x4096 pixels (depending on
your SEM)
=B7 Square pixels in any mode
=B7 Connects to any existing SEM or STEM
=B7 100% reliable for your SEM and lowest noise level due to unique, optical
isolation on board
=B7 Grabs images in slow scan and full photo resolution
=B7 Can grab 2 simultaneous images (for example SE and BSE fields)
=B7 Unique, programmable oversampling factor dramatically reduces image=
noise
in real time
=B7 The grabbed images are immediately available to the user in the PC RAM
=B7 They are stored in full resolution, 256 grey levels per pixel (frame
integration uses 16 bits per pixel during grabbing)
=B7 Easy transfer to any other application (OLE 2 or clipboard)
=B7 Distance measurement, extraction, filtering, line / rectangle drawings,
line scan function
=B7 Selectable, enhanced image compression engine reduces image size on disk
by 90% with little image degradation
=B7 Full compatible with Windows 3.1x, Windows 95 and Windows NT.

Options
=B7 EDX mapping grabbing mode mixes elemental maps with SE or BSE image
=B7 Ultra high resolution photo replay allows image to be photographed later
from disk
=B7 HP Laserjet enhancement card allows photographic quality in 10 seconds
=B7 Powerful yet simple programming language makes the user=92s job easier



E.L.I. sprl (Belgium)

Technical services Sales department =09
Rue Lossignol, 51 Avenue des Jardins, 46
1401 - Baulers 1030 - Bruxelles
tel: (32) 67 21 25 07 tel: (32) 2 726 31 02
fax: (32) 67 22 09 53 fax: (32) 2 726 08 65
Email: jleclef-at-hypercon.com Email: orion-at-infoboard.be

See also our WEB site http://www.microscopy-uk.org.uk


At 16:33 24/03/1997 -0500, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20


Best regards,



Paul Vanderlinden.
Sales Manager.

=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D
See our web site: http://www.microscopy-uk.org.uk =20

To contact us:

E.L.I. sprl

Technical support:=20
Jean-Louis Leclef: Phone: +32 67 21 25 07 =20
Fax : +32 67 22 09 53=20
Email: jleclef-at-hypercon.com
Sales support:
Paul Vanderlinden: Phone: +32 2 726 31 02 =20
Fax : +32 2 726 08 65
Email: orion-at-infoboard.be=20
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D
=
=20
=20





From: Ya Chen :      ychen14-at-facstaff.wisc.edu
Date: Thu, 27 Mar 1997 08:46:08 -0600
Subject: Re: teflon o-rings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

}
} Does anyone know of a source for teflon o-rings with a thickness of {100 uM?
}
} Thanks,
} Sandy Simon
}
} Sanford M. Simon
} Laboratory of Cellular Biophysics
} Box 304
} Rockefeller University
} 1230 York Avenue
} New York, N.Y. 10021
} 212-327-8130 (voice)
} 212-327-8022 (fax)
} simon-at-rockvax.rockefeller.edu (e-mail)


We purchased teflon o-ring 9 years ago from Bal Seal Engineerig Co., 620 W.
Warner Ave. Santa Ana, CA 92707-3398. Tel: 714-557-5192.

Ya Chen





Ya Chen

=========================================================================
\ / Integrated Microscopy Resource (IMR)--
\ / __ an NIH Biomedical Research Resource TEL : 608-263-8481
\/ / / University of Wisconsin-Madison FAX : 608-265-4076
/ / / 1675 Observatory Drive #159 Email1:ychen14-at-facstaff.wisc.edu
/ /__/_ Madison, WI 53706 Email2:chen-at-calshp.cals.wisc.edu
=========================================================================
IMR WWW Home Page: http://www.bocklabs.wisc.edu/imr.html






From: Mohan Kalyanaraman :      mxkalyan-at-pau.mobil.com
Date: Thu, 27 Mar 97 09:18:17 EST
Subject: Help in locating Practical e- microscopy book by Edington

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear All,

I request your help in locating a publisher or bookseller who carries
the book "Practical electron microscopy in Material Science" by
J.W.Edington. It was originally published as a 4 vol. series by
Philips. Van Nostrand who published the combined volume in 1976 say
the book is out of print.

Thanks for your help.

Mohan Kalyanaraman
Sr. Staff Material Scientist
Mobil Technology Company
Paulsboro, NJ 08066
609-224-3989





From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Thu, 27 Mar 1997 10:18:02 -0400
Subject: RE:ElecChanneling Contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Chapter 3 of the book "Advanced Scanning Electron Microscopy & X-ray Micro
Analysis" by Newberry, Joy, et. al, Plenum Press 1986 is entitled "Electron
Channeling Contrast in the SEM", and ought to provide the information you
are looking for.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Thu, 27 Mar 1997 10:19:24 -0400
Subject: Addr for Bob Baier

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Bob Baier at the Univ of Buffalo apparently saw my name on this list, and
sent me a personal message about a week ago. I have since tried several
times to reply to him using the address: baier-at-ubvms.cc.buffalo.edu -
(which I copied directly from the original message I received from him),
but keep getting rejections due to 'local delivery' problems. Does anyone
happen to know a better address for him, or have suggestions on how to get
through to him with this one?

TIA

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: SEMTRADER-at-aol.com
Date: Thu, 27 Mar 1997 11:22:05 -0500 (EST)
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

There are a few vendor that offer Image capturing for the SEM. Most of the
system we have installed on our refurbished SEM are supplied by Evex
Analytical. There number is 609-252-9192.


Drew




From: Gary Chinga :      gary-at-nvg.ntnu.no
Date: Thu, 27 Mar 1997 17:43:14 +0100 (MET)
Subject: Microscopes...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hi!

I am trying to find out what is the resolution of a Confocal Scanning
Laser Microscope. Does it depend on the type of laser used?

I also have another question concerning SEM microscopes. Why is the
column of the SEM shorter than the TEM. It is because of a simpler
electron optic system in the SEM?.

I know that these are basic questions, but answers to questions like
these are impossible to find in textbooks

Thanks.

GCH.





From: colijn.1-at-osu.edu (Henk Colijn)
Date: Thu, 27 Mar 1997 11:49:03 -0500
Subject: Re: Help in locating Practical e- microscopy book by Edington

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Edington has been available in an Indian reprint from

Tech Books
4012 Williamsburg Ct.
Fairfax, VA 22032
703 352-0001
703 352-8862 FAX



} Dear All,
}
} I request your help in locating a publisher or bookseller who carries
} the book "Practical electron microscopy in Material Science" by
} J.W.Edington. It was originally published as a 4 vol. series by
} Philips. Van Nostrand who published the combined volume in 1976 say
} the book is out of print.
}
} Thanks for your help.
}
} Mohan Kalyanaraman
} Sr. Staff Material Scientist
} Mobil Technology Company
} Paulsboro, NJ 08066
} 609-224-3989

Henk Colijn colijn.1-at-osu.edu OSU Campus Electron Optics Facility
-------------------------------------------------------------------
Pereant qui ante nos nostra dexerunt.






From: Woody.N.White-at-mcdermott.com
Date: 3/25/97 2:27 PM
Subject: Electron Channeling contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hope I am not double posting... The orginal didn't show up and may
have been lost is cyberspace...??? _W_



It is not uncommon for me to employ electron channelling contrast.
Almost all of my experience is with stainless steels, nickel
superalloys (Inconels), Zirconium alloys, and uranium compounds. The
ecc is a weak signal and is not always easy to produce.

A few ideas....

Use BSE imaging with large beam currents and high BSE signal gain
(contrast). It may be necessary let any response from areas of higher
or lower Z than the matrix of interest go saturated black and/or white
to achieve this.

Use a "normal" incident beam (0 degrees tilt)

Contrary to some advice I have recieved, I find that lower beam
voltages (10 kv) work better than higher (20-30 kV). I have no proof,
but suspect that the lower penetration depth images the surface grains
without "confusing and diluting" the image with BSE returns from
sub-surface grains with different orientations.

Surface preparation is VERY important. A very well polished surface, free
from
surface damage is required or the signal will be obscured. Some materials
are
easier than others to prepare. On occation, I have had to send samples back
to
our met-lab several times before a satisfactory surface is available. A
very
light "attack" polish (not really an etch) can be helpful during final
polish
for some materials. I have wanted to try a light ion beam cleaning, but
don't
have one.

Incident beam angle (with specimen at 90 nominal) changes resulting from the

raster can strongly affect contrast. Experiment with working distance vs.
magnification to see what works best for you. Along this line... Don't
expect
to be able to generate multiple image mosaics which match well. A grain
that is
dark on one edge of an image may will be light when the stage is translated
one
frame over because the incident angle is nolonger the same.


Hope this helps...

Woody White

http://www.geocities.com/capecanaveral/3722

BTW something happened to my email pgm half way through??? Hope this
dosen't
"run over" line length too much!

______________________________ Reply Separator
_________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Good afternoon everyone.

Can anyone refer me to a basic reference on electron channeling
contrast,
or failing that, give me a few hints on optimal conditions for observing
this type of contrast ?

TIA

Glenn

Glenn Poirier Phone (514) 398 6774
Electron Microprobe Laboratory Fax (514) 398 4680
Earth and planetary Science
Mcgill University

THERE ARE THREE SIDES TO EVERY STORY:
yOURS, MINE AND THE TRUTH




From: Ya Chen :      ychen14-at-facstaff.wisc.edu
Date: Thu, 27 Mar 1997 12:46:49 -0600
Subject: Re: Microscopes...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v03007801af6070cff582-at-[144.92.238.41]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

} Hi!
} ........

} } I also have another question concerning SEM microscopes. Why is the
} column of the SEM shorter than the TEM. It is because of a simpler
} electron optic system in the SEM?.
}
} I know that these are basic questions, but answers to questions like
} these are impossible to find in textbooks
}
} Thanks.
}
} GCH.


Hi Gary,

Basicly, a TEM has condenser, objective, intermediate, and projection
lenses, plus a fluorescent screen and a film chamber. In contrast, SEM has
only condenser and objective lenses, plus a speciemen chamber. That is
obviously why SEM has a short column.


Ya Chen


Ya Chen

=========================================================================
\ / Integrated Microscopy Resource (IMR)--
\ / __ an NIH Biomedical Research Resource TEL : 608-263-8481
\/ / / University of Wisconsin-Madison FAX : 608-265-4076
/ / / 1675 Observatory Drive #159 Email1:ychen14-at-facstaff.wisc.edu
/ /__/_ Madison, WI 53706 Email2:chen-at-calshp.cals.wisc.edu
=========================================================================
IMR WWW Home Page: http://www.bocklabs.wisc.edu/imr.html

The Integrated Microscopy Resource and Carnegie Mellon University will
be sponsoring a symposium and short course on multi-photon excitation
imaging, August 9-10, 1997, in Cleveland Ohio.






From: paulc-at-gps.caltech.edu (Paul K. Carpenter)
Date: Thu, 27 Mar 1997 11:03:59 -0800
Subject: Re: Peak Overlaps (Ti Kb on V Ka)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Ritchie Sims wrote,

} I ask because I concluded recently that I couldn't, with my EDS-only
} system, analyse for V in titanomagnetites, and that it probably
} wasn't possible even with WDS.
} If I ask my system to look for V in my standard Rutile (TiO2), it
} always finds a few tenths of a %, even though my software
} (LINK ZAF4/FLS) strips out Ti Kb along with the Ka. I have no way of
} telling whether this V is genuine or not, but I suspect that it
} isn't, as my supposedly-pure Ti metal standard also gets credited
} with a small amount of V, which I doubt.
} I was recently given another rutile standard, and, blow me down, it
} is reputed to have 0.4% V by WDS analysis.
} Anyone got any feeling for whether this is likely to be true?

The general case for interference with the transition elements is that the
Z-1 element Kb peak overlaps the Z element Ka peak. Ti Kb on V Ka is what
you bring up.

For EDS analysis, assuming that linear least squares deconvolution is used,
the fit for Ti should handle both Ti Ka and Kb if they are grouped together
as your EDS reference. Any residual could then appear as a positive value
for V. This may reflect a calibration difference between your sample and
reference spectra (I'm assuming we are not talking standardless EDS here).

For WDS measurement one has the choice of PET or LIF; the latter has better
resolution and should be used to reduce the magnitude of the overlap to
begin with. I find that (using LIF) the measured V Ka k-ratio on synthetic
(pure) TiO2 is typically less than 0.5%, which compares with:

} I was recently given another rutile standard, and, blow me down, it
} is reputed to have 0.4% V by WDS analysis.

My guess is that the analyst did not correct for the overlap. In the
absence of any more sophisticated software, you can make a correction for a
simple overlap like this via:

corrected V K-ratio = measured V K-ratio - AK * measured Ti K-ratio

where AK is the apparent V K-ratio measured on pure TiO2. Thus, when
analyzing TiO2 the resulting V concentration is zero, and when analyzing a
phase containing no Ti the correction is zero. This method corrects the
K-ratio before passing to the ZAF program, but you could just as well make
a small correction like this on the weight percent data. In practice I
find that this linear correction reduces the apparent V in pure TiO2 by an
order of magnitude, i.e. from 0.X% to 0.0X%, so especially several hundred
ppm V in rutile should be viewed with scepticism.

You can also do a wavelength scan on your rutile looking for the presence
of the V *Kb* peak, since this is not overlapped. If you see this peak,
then you know there is V in the sample.

This is not a bad overlap. Consider Pb La on As Ka using LIF. This is a
total overlap with no real possible solution via this linear correction.
The solution in a case like this is to use As Kb and either Pb Lb or Pb Ma
as your lines (the application is for sulfides or arsenides). As long as
there are no absorption edges between the Ka and Kb lines, or La and Lb
lines, the mass absorption coeffiecients are usable; one really has no
other choice.

Have fun,

Paul


+----------------------------------------------------+
| Paul K. Carpenter paulc-at-gps.caltech.edu |
| Division Analytical Facility |
| Geological and Planetary Sciences MC 100-23 |
| California Institute of Technology |
| Pasadena, CA 91125 |
| 818-395-6126 (X-ray Lab) 818-568-0935 (Dept. FAX) |
+----------------------------------------------------+






From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 27 Mar 1997 14:33:26 -0500 (EST)
Subject: Re: Nitrogen Drying

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear Chris,

} I have just read an artical that includes in its protocol for fluorescent
} dyeing, blot drying of the samples, then drying under a stream of nitrogen.
}
} My question was what is the benefit of drying under a stream of nitrogen as
} opposed to air drying? My first thought was perhaps to reduce any oxidation.

Oxidation is one consideration. The lack of water vapor in the N2
stream is another, and lack of dust, oil droplets, etc. in N2--these are
often present in compressed air--is a third.

} Can anyone shed any light on this matter.
}
If I did, would you just send it back longer? ;-)
Yours,
Bill Tivol




From: Cox, Tom :      Cox#m#_Tom-at-msmail.msd.lmsc.lockheed.com
Date: 27 Mar 1997 10:55:15 -0800
Subject: Need some info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

My LEO service engineer recommended that I contact you to get an e-mail address
for Chris Martinic at the University of New South Wales. Chris has information
about image grabbers that I need. Please send me his e-mail address. My address
is:

tom.cox-at-lmco.com

Thank you,

Thomas J. Cox
Lockheed Martin Missiles & Space Co.




From: John Fournelle :      johnf-at-geology.wisc.edu
Date: Thu, 27 Mar 1997 13:49:57 -0600
Subject: e beam + epoxy = ??

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {v03007806af607ed3d42c-at-[144.92.137.115]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

I would like to know what goes on inside the chamber of my electron microprobe
when the focused, moderate to high current (10 - 200 nA) electron beam hits
epoxy. Visually, a crater forms. What are the solid/gaseous by-products of
that reaction? Any ideas, educated guesses, and/or experimental or
theoretical results (or suggestions where to look) would be appreciated.
Thanks.

John

John Fournelle
Electron Microprobe Lab office: (608) 262-7964
Dept of Geology & Geophysics fax: (608) 262-0693
University of Wisconsin home: (608) 274-2245
1215 West Dayton St. email: johnf-at-geology.wisc.edu
Madison, WI 53706 amateur radio: WA3BTA/9
Personal http://geology.wisc.edu/~johnf/
Probe lab http://geology.wisc.edu/~johnf/sx51.html

"The first rule of all intelligent tinkering is to save every cog and wheel."
Aldo Leopold






From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 27 Mar 1997 14:58:19 -0500 (EST)
Subject: Re: Microscopes...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} I am trying to find out what is the resolution of a Confocal Scanning
} Laser Microscope. Does it depend on the type of laser used?
}
Dear Gary,
The resolution of a CSLM is ~1/3 to ~1/2 the wavelength of the
ilumination in the x-y plane and about 2x worse in the z-direction.
Jim Pawley's Handbook of Confocal Microscopy has a very good description
of the theory. (I hope I got the title right--the book is at my desk &
I'm not.) It will depend on the wavelength of light used, but not on
whether the laser is pulsed or continuous wave, so whether it depends
on the "type of laser used" depends on what you mean. There may be
CSLM's which use frequency-doubled light, so a pulsed laser is neces-
sary to provide sufficient intensity to get the non-linear optics to
work, so if this is what you mean, then whether the CSLM works at all
can depend on the type of laser used, but once you get light of a par-
ticular wavelength, the resolution won't depend on where that light came
from. Since I'm not an expert on this, you should probably check the book.
Yours,
Bill Tivol




From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Thu, 27 Mar 1997 10:10:33 -1000 (HST)
Subject: Re: Looking for veteran SEM service engineers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} } Hello Microscopists,
} }
} } Can anyone suggest names of service engineers with experience on
} } vintage Cambridge electron microscopes? I have a 1971 Cambridge
} } Stereoscan S4 SEM in need of service. I'd like to locate an
} } experienced technician in the metro Atlanta area or the southeastern
} } US, but I would appreciate all references and information. Thanks,

Clark Houghton at Secondary Images is a real wizard with the old
Cambridges. He is in Ohio (but was willing to come all the way to
Hawaii... what a hardship!) at (513) 927-5373.

Tina

http://www.pbrc.hawaii.edu/bemf/microangela
****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************





From: Nancy.P.Piatczyc-at-williams.edu
Date: Thu, 27 Mar 1997 15:10:10 -0500
Subject: Re:looking for veteran SEM service

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Hello! We have a Cambridge Stereoscan and for service use Scanners Corporation
in Eldersburg, MD - tel. 1-410-549-3800 I believe they cover the East Coast
and possibly a bit further out as well. They take great care of our SEM.
Nancy






From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Thu, 27 Mar 1997 16:16:18 -0400
Subject: RE: SEM ColumSize

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

In response to the question as to why SEM colums are so much smaller than
those of TEMs, I offer the following:

The colums of SEMs basically contain only two lenses: a 'first' lens (often
incorrectly called the condenser lens) which produces demagnification of
the spot from the electron gun, and a 'second' lens (usually incorrectly
called the objective lens) which is used to focus the spot image produced
by the first lens onto the specimen forming the electron probe which
generates the signals that are emitted from the sample. Sometimes the
first lens is a compound lens, actually giving a total of three lenses;
nonetheless, they function as a two-lens system overall.

This entire electron optical system of an SEM corresponds in general
character and function to the 'condenser lens system' (i.e. the system of
lenses that is above the specimen chamber) of a TEM. That is, the
condenser lens system of a TEM is basically a two-lens system which
functions to control the illumination that strikes the specomen (just as
the two lens system of the SEM does).

In a TEM; however,there is an additional image-forming lens system below
the specimen consisting of the objective lens, which produces the primary
electron image from the specimen, and then a series of from 3 to 6
additional lenses which produce the wide range of magnifications we expect
from TEMs now-a-days, plus providing the variety of additional imaging
functions commonly available (selected area diffraction, convergent beam
diffraction, etc.). No lens is directly involved in the image-forming
process in an SEM, and so these additional lenses are not needed.

Thus there are three or four times as many lenses in a TEM as there are in
an SEM. In addition, whereas most TEMs are designed to operate at electron
accelerating voltages from 80 kV to several hundred kilovolts, most SEMs do
not use accelerating voltages much above 30 kV. Since it is much easier to
deflect these lower energy electrons, the magnetic fields in SEM lenses do
not need to be as strong as those in TEMs; consequently they require fewer
turns of wire in their excitation coils, and so can be physically much
smaller.

I hope this answers your question satisfactorily. If not, don't hesitate
to let me know. I can send you column diagrams which illustrate the above
points, if needed.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Thu, 27 Mar 1997 16:16:18 -0400
Subject: RE: SEM ColumSize

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

In response to the question as to why SEM colums are so much smaller than
those of TEMs, I offer the following:

The colums of SEMs basically contain only two lenses: a 'first' lens (often
incorrectly called the condenser lens) which produces demagnification of
the spot from the electron gun, and a 'second' lens (usually incorrectly
called the objective lens) which is used to focus the spot image produced
by the first lens onto the specimen forming the electron probe which
generates the signals that are emitted from the sample. Sometimes the
first lens is a compound lens, actually giving a total of three lenses;
nonetheless, they function as a two-lens system overall.

This entire electron optical system of an SEM corresponds in general
character and function to the 'condenser lens system' (i.e. the system of
lenses that is above the specimen chamber) of a TEM. That is, the
condenser lens system of a TEM is basically a two-lens system which
functions to control the illumination that strikes the specomen (just as
the two lens system of the SEM does).

In a TEM; however,there is an additional image-forming lens system below
the specimen consisting of the objective lens, which produces the primary
electron image from the specimen, and then a series of from 3 to 6
additional lenses which produce the wide range of magnifications we expect
from TEMs now-a-days, plus providing the variety of additional imaging
functions commonly available (selected area diffraction, convergent beam
diffraction, etc.). No lens is directly involved in the image-forming
process in an SEM, and so these additional lenses are not needed.

Thus there are three or four times as many lenses in a TEM as there are in
an SEM. In addition, whereas most TEMs are designed to operate at electron
accelerating voltages from 80 kV to several hundred kilovolts, most SEMs do
not use accelerating voltages much above 30 kV. Since it is much easier to
deflect these lower energy electrons, the magnetic fields in SEM lenses do
not need to be as strong as those in TEMs; consequently they require fewer
turns of wire in their excitation coils, and so can be physically much
smaller.

I hope this answers your question satisfactorily. If not, don't hesitate
to let me know. I can send you column diagrams which illustrate the above
points, if needed.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:313-763-4788; Ph:313-764-3321






From: Larry Stoter :      LPS-at-teknesis.demon.co.uk
Date: Thu, 27 Mar 1997 21:00:20 +0000
Subject: Re: Microscopes...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Not my field, but wavelength of illumination will be one determining factor.

} I also have another question concerning SEM microscopes. Why is the
} column of the SEM shorter than the TEM. It is because of a simpler
} electron optic system in the SEM?.

Two answers:

1. Voltage - generally, SEMs work at a maximum of 30kV while TEMs typically
work from a minimum of 100kV. Higher voltage electron have much greater
kinetic energy and are thus more diffcult to deflect. Even with somewhat
stronger lenses, the additional energy of the electron requires a little
more distance to have the required effect.

2. More significantly, an SEM column is only equivalent to the part of the
TEM colum ABOVE the specimen. In a TEM, additional lenses are needed after
the specimen to focus the beam into an image. In a SEM, the electron beam
is aleady focused to a probe at the specimen, and only needs 'intensity'
detectors. Indeed, if you do secondary electron imaging on a bulk specimen
in a TEM/STEM type if instrument, the bottom half of the column is not used.

} I know that these are basic questions, but answers to questions like
} these are impossible to find in textbooks
}
} Thanks.
}
} GCH.

Regards,


---------------------------------------------------------------
Dr L. P. Stoter Technical Editor, MICROSCOPY & ANALYSIS
Technesis
17, Rocks Park Road email: LPS-at-teknesis.demon.co.uk
Uckfield, E. Sussex Phone: +44 (0)1825 766911
TN22 2AT Fax: +44 (0)1825 766911
United Kingdom
---------------------------------------------------------------






From: Raymond F Egerton :      egerton-at-phys.ualberta.ca
Date: Thu, 27 Mar 1997 14:37:28 -0700 (MST)
Subject: Re: searching for image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Seth: we built an SEM image-capture system based on a National
Instruments (AT-MIO-16E-10) data-acquisition card, controlled by a
program written in Visual Basic. Hardware cost was about $US 3000
including Pentium computer and 1200 dpi laser printer. Brief details of
the system will be presented at the June MSC conference, 2-page abstract
appearing soon on the conference web page: http://www.ualberta.ca/~mmid/msc

Ray Egerton, Physics Dept, University of Alberta, Edmonton, Canada T6G 2J1
Phone: 403-492-5095, FAX: 403-492-0714, e-mail: egerton-at-phys.ualberta.ca
------------------------------------------------------------------------




From: Ronald Lee Austin :      rla-at-mindspring.com
Date: Thu, 27 Mar 1997 18:48:59 -0500
Subject: The lenght of an SEM column

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I have not had much experience with scanners, but it seem to me the
answer
to why there columns are not as long as the TEM's are maybe in the fact
that you do not have the objective len and stage or the projector or
diffraction lens to deal with. Thats my best guess. As for Confocal
Scanning I have had no experience at all with this device. sorry!

Ron Austin
rla-at-mindspring.com






From: Mary Mager :      mager-at-unixg.ubc.ca
Date: Thu, 27 Mar 1997 20:59:14 -0800
Subject: Re: e beam + epoxy = ??

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Dear John,
When I did a careful study of SEM sample contamination, using a polished
copper sample and carefully regulated kV, current and magnification, the
deciding factor seemed to be how recently someone had hit epoxy with the
beam. One "hit" seemed to result in much worse contamination for about three
days. In high current situations you can see the epoxy "boil". I think the
vapourised hydrocarbons fly about your chamber for quite a while. Cleaning
with a N2 purge, like the SEMclean system (I use a home-built one), does
seem to help.
I seem to recall that someone did a study of contamination, air-jets and
such in Microbeam Analysis in the early '80s. May have been Bastin.
You wrote:

} I would like to know what goes on inside the chamber of my electron microprobe
} when the focused, moderate to high current (10 - 200 nA) electron beam hits
} epoxy. Visually, a crater forms. What are the solid/gaseous by-products of
} that reaction? Any ideas, educated guesses, and/or experimental or
} theoretical results (or suggestions where to look) would be appreciated.
} Thanks.
}
} John
Hope this helps,
Mary
Mary Mager
Electron Microscopist
Metals and Materials Eng., UBC
6350 Stores Rd.
Vancouver, B.C. V6T 1Z4
CANADA
tel:604-822-5648, fax:604-822-3619
e-mail: mager-at-unixg.ubc.ca





From: Glen MacDonald :      glenmac-at-u.washington.edu
Date: Thu, 27 Mar 1997 14:43:26 -0800
Subject: re:image capture system

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Seth,

You might want to check out 4Pi Analysis. they have a a Mac-based
image capture system for SEMs that is priced very competitively. There
boards and software offer total control over the SEM, with image
acquistion by NIH Image, Photoshop, IPLab Spectrum, etc.

4pi Analysis
3500 Westgate Dr., Suite 403
Durham, NC 27707
(919) 489-1757

Regards,
Glen MacDonald
Virginia Bloedel Hearing Research Center
Box 35-7923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156
glenmac-at-u.washington.edu

/*----------------------------------------------------------------------
-*/
/ The box said "Requires Windows 95 or better.", so I bought a
Macintosh. /
/*----------------------------------------------------------------------
-*/





From: jeanne_barker-at-merck.com (Jeanne Barker)
Date: Fri, 28 Mar 1997 08:10:27 -0500
Subject: None

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




From: Michael OKeefe :      Michael_OKeefe-at-macmail.lbl.gov
Date: 28 Mar 1997 09:49:55 -0700
Subject: In Memoriam Ñ Mark Fendorf

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Subject: Time:9:50 AM
OFFICE MEMO In Memoriam - Mark Fendorf Date:3/28/97

In Memoriam - Mark Fendorf

It is with great sadness that we at NCEM announce the death of Mark Fendorf.


Mark John Fendorf died on March 14, 1997 at the age of 36. As a
collaborative post-doctoral scientist at LBNL's National Center for
Electron Microscopy, Mark played a central role in NCEM's outreach to
external facility users. He was an outstanding microscopist, a resourceful
materials scientist and a delightful colleague to all members of NCEM. In
his role of resident expert microscopist he collaborated with many NCEM
users and helped countless visitors of the facility. He earned his Ph.D. in
materials science from UC Berkeley in 1992. His scientific work focused
on electron beam micro-characterization of a number of different
materials, including high-temperature superconductors, catalysts,
fullerines and most recently the adsorption of heavy metals on soil
minerals. He was an active member of several scientific societies,
including the Microscopy Society of America and the American Ceramic
Society. In 1991, he co-founded the UC Berkeley chapter of the Materials
Research Society. Outside of LBNL, he was a dedicated and enthusiastic
volunteer for the Boy Scouts of America where he was a leader for 16
years, served on the Monterey Bay Area Council, and acted as Director of
Troop Leadership Training for a number of years.

During his progressively debilitating illness, Mark continued his work at
NCEM with remarkable dedication and tenacity. His strength of will and
his determination to overcome his handicap caused by the illness were
admired by all those who knew him. Mark's untimely death has left the
Center permanently diminished. He will be greatly missed by his NCEM
colleagues and the many users and visiting scientists at the facility.


Mark Fendorf is survived by his parents, Ken and Virginia Fendorf and his
brothers Scott and Dale. In recognition of his sustaining enthusiasm for
the science of electron microscopy, and in an effort to help others
continue where Mark had to leave off, his family has established a
memorial fund in support of NCEM's outreach program.

Donations to the Mark Fendorf memorial fund can be made to:
World Savings, Aptos Branch, Mark Fendorf Memorial Fund, 7970 Soquel
Drive, Aptos, CA 95003







From: William Tivol :      tivol-at-wadsworth.org
Date: Fri, 28 Mar 1997 13:55:48 -0500 (EST)
Subject: Re: e beam + epoxy = ??

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear John,

} I would like to know what goes on inside the chamber of my electron microprobe
} when the focused, moderate to high current (10 - 200 nA) electron beam hits
} epoxy.

When an electron passes through matter, the energy of the electron
is transferred to the material in the form of ionizations and excitations.
About 30 eV is transferred for each ionization, and most of the energy
transferred goes into primary or secondary ionizations. The products, ob-
viously far from thermodynamic equilibrium, react with other components
of the material causing chemical bond breaking and free radical formation.
Finally (a few microseconds later) these reactive species form more stable
products, which are often small organic molecules in the case of electrons
incident on epoxy resin. There is a statistical distribution of products
which depends on the nature of the resin; since many of these are volatile,
they will travel throughout the chamber (they also gain some kinetic energy
from the energy transferred from the initial electron).

} Visually, a crater forms.

This should make sense in light of the above.

} What are the solid/gaseous by-products of
} that reaction? Any ideas, educated guesses,

The above are educated guesses...

} and/or experimental or
} theoretical results

and theoretical results.

} (or suggestions where to look) would be appreciated.

My info comes from Friedlander, et al., Nuclear and Radiochemistry;
I do *not* recommend this book. A good book on radiation chemistry would
be a good place to start, but I don't know any titles.
Yours,
Bill Tivol




From: bozzola-at-siu.edu (John J. Bozzola) (by way of Nestor J. Zaluzec)
Date: Sun, 30 Mar 1997 22:03:21 -0500
Subject: SEM Jeol high res standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Jeol sells a high resolution standard (gold on carbon) that we would like
to purchase. Does anyone have info: (part number, price, where to order)?
Many thanks.

####################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Neckers Building, Room 146 - B Wing
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################






From: A Wilson :      awilson-at-aw.u-net.com
Date: Sun, 30 Mar 1997 10:50:01 +0100
Subject: PROPOSED NEW IMMUNOCYTOCHEMISTRY USENET NEWSGROUP

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

A formal proposal to create a new group tentatively called
sci.bio.immunocytochem was posted to news.announce.newgroups
on 17.3.97. This is a reposting of that proposal. Please
have a read and let me know what you think.

REQUEST FOR DISCUSSION (RFD)
unmoderated group sci.bio.immunocytochem

This is the 2nd Request For Discussion (RFD) for the creation of a
world-wide unmoderated Usenet newsgroup sci.bio.immunocytochem,
currently being discussed in news.groups

Suggestions for improvements to this proposal are welcome. Discussion
about it should take place in news.groups. A vote is expected to be
held in about three to four weeks.

This is not a Call for Votes (CFV); you cannot vote at this time.
Procedural details are below.

CHANGES from previous RFD:

2nd RFD posted because more than 60 days have elapsed since 1st RFD.
There are (minor changes in Distribution and Newsgroup line.

Newsgroup line:
sci.bio.immunocytochem Immuno-labelling of biological material.

RATIONALE: sci.bio.immunocytochem

Immunohistochemists and immunocytochemists already enjoy the benefits
of online communication, utilizing e-mail, accessing web sites, and
subscribing to specialised mailing lists. Usenet newsgroups are also
popular, but this is less obvious because articles with
immunocytochemical/immunohistochemical content get posted to many
different newsgroups. Most articles are posted to a favourite five or
six newsgroups including bionet.cellbiol, sci.med.immunology and
sci.techniques.microscopy, but often articles get posted to any one of
fourteen or fifteen newsgroups in the sci. and bionet. heirarchies.
Some of these are listed in the distribution list at the end of this
proposal.

In my view, no existing newsgroup fulfils the criteria necessary to
attract all the various immunocytochemistry postings. I do not wish
to draw users away from other newsgroups, only to encourage scientists
to share their knowledge and expertise on immunocytochemistry in the
most effective manner. In response to my proposal to create a
newsgroup dedicated to the discussion of immunocytochemistry and
immunohistochemistry, I have received e-mail and faxes from
researchers all over the world offering their support and
encouragment.

Immunocytochemistry and immunohistochemistry are not subdivisions of
immunology, molecular biology or chemistry. Microscopy, although
essential, is only a small part of the story. Immunocytochemistry and
immunohistochemistry are multi- disciplinary, therefore discussions
are destined to stay distributed amongst the different newsgroups
until they are all brought together under one umbrella. This would
then act as a focus point for all the immunocytochemists who are
already Internet users, and encourage new subscribers to Usenet.

CHARTER: sci.bio.immunocytochem

This is a newsgroup for the exchange of information relating to
immunocytochemistry and immunohistochemistry. This unique research
tool is used to locate and identify specific molecules in biological
material, at the microscopical level.

Articles posted to this group must be relevant to one or more aspects
of the above. The kind of subjects that may be discussed include
techniques, theory, presentation of results, requests for
collaboration, history, equipment, publication references, notice of
events, tips and trouble-shooting, jobs offered andwanted, jokes,
stories and new ideas, so long as the posting bears a direct relevance
to the central theme. There will be a list of Frequently Asked
Questions (FAQs) to help newcomers.

A relevant posting could just be a simple question or answer, for
example "Has anyone got any experience with this reagent ?"or "Which
course could I attend to learn more about immunogold labelling?".
There will be articles reminding people to read the list of FAQs prior
to posting their own article. Usenet readers may get involved in
complex discussions about, for example, multiple labelling, proper use
of control experiments, microwave antigen retrieval or quantitative
measurements. Remember that articles posted to a newsgroup are
intended for a wide readership, so if you have information which
concerns only one or two people then please don't use this newsgroup,
use e-mail.

Commercial advertisements for services, equipment or reagents violate
the charter unless one or more of the following apply: (a)The
advertisement is part of a comprehensive article designed specifically
to address issues raised in earlier articles posted to the group (b)A
general reference to the type of product does not suffice for
technical reasons and it is necessary to specify the exact commercial
product (c) The information is offered primarily for the benefit of
the readers (d)The advertisement is for second-hand equipment specific
to immunocytochemistry (e) Requests or offers for free products are
acceptable if they are not part of a sales promotion.

END CHARTER.

PROCEDURE:

This is a request for discussion, not a call for votes. In this phase
of the process, any potential problems with the proposed newsgroups
should be raised and resolved. The discussion period will continue
for a minimum of 21 days (starting from when the 2nd RFD for this
proposal is posted to news.announce.newgroups), after which a Call For
Votes (CFV) may be posted by a neutral vote taker if the discussion
warrants it. Please do not attempt to vote until this happens.

All discussion of this proposal should be posted to news.groups. This
RFD attempts to comply fully with the Usenet newsgroup creation
guidelines outlined in "How to Create a New Usenet Newsgroup" and "How
to Format and Submit a New Group Proposal". Please refer to these
documents (available in news.announce.newgroups) if you have any
questions about the process.

DISTRIBUTION:

This RFD has been posted to the following newsgroups:
news.announce.newgroups,news.groups,bionet.cellbiol,
bionet.diagnostics,bionet.immunology,bionet.microbiology,
bionet.molbio.methds-reagnts,sci.bio.misc,sci.med.immunology,
sci.techniques.microscopy

This RFD will be reposted to the following newsgroups after its
posting in news.announce.newgroups:
bionet.molbio.proteins,bionet.neuroscience,bionet.plants,
sci.bio.microbiology,sci.med,sci.med.laboratory,sci.misc,
sci.nanotech

This RFD will also be reposted to the following mailing lists after
its posting in news.announce.newgroups:

Histonet mailing list: {histonet-at-pathology.swmed.edu}
Information pertaining to the technical aspects of histology and
histopathology such as tissue fixatives and processing, routine
histology, special stains, immunohistochemistry, in-situ
hybridization etc. To subscribe type "subscribe digest" into the
subject box and leave the text box empty, or to subscribe to the full
service just type "subscribe". For more info access web site
http://www.mwrn.com/subject/histonet.htms

Microscopy Society of America listserver:
Questions/comments/answers in the various fields of Microscopy
Currently over 3000 subscribers. To subscribe send the message
"subscribe" to {Listserver-at-MSA.Microscopy.Com}
then send messages in plain text to
{Microscopy-at-MSA.Microscopy.Com}
For more info access web site http://www.amc.anl.gov/
Docs/anl/Nestor/Software/telecommList.html

Stanford University list server
To subscribe, send a message to
{majordomo-at-pathology.stanford.edu} with "subscribe ipox-l" in
the body of your message. This list helps pathologists and other
laboratory professionals to exchange information about
immunoperoxidase methods.

This RFD will also be reposted to the following web-sites after its
posting in news.announce.newgroups:

Royal Microscope Society
http://www.rms.org.uk
Web Master Dr R. A. D. Mackenzie
{r.a.mackenzie-at-open.ac.uk}

Center for Cell Imaging Department of Cell Biology
Yale University School of Medicine
Introduction to Immunocytochemistry
http://info.med.yale.edu/cellimg/CCIimmuno.html
Web Master Paul Webster
{ paul_webster-at-yale.edu}

Proponent: Amanda Wilson {awilson-at-aw.u-net.com}
Proponent: Paul Monaghan { monaghan-at-icr.ac.uk}
Mentor: Jonathan Grobe {grobe-at-netins.net}












From: Simon Watkins :      swatkins-at-pitt.edu
Date: Sat, 29 Mar 97 14:01:29 -0500
Subject: balzers e-beam gun

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Simon Watkins * EMC.Ver #2.5.02 ] --

Hi folks:

simple question, when recharging a balzers gun with platinum, where should
the tip of the carbon rod holding the platinum pellet be with repect to the
tungsten coil?

Tx

simon

--
-------------------------------------------------
Simon C. Watkins Ph.D
Associate Professor
Director Center for Biologic Imaging
University of Pittsburgh
Pittsburgh PA 15261
tel 412-648-3051
fax 412-648-2004
-----------------------------------------------




From: kennedy-at-nsi.edu (Grace Kennedy)
Date: Fri, 28 Mar 1997 17:21:30 -0800
Subject: UV polymerizaton

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I'm about to attempt UV polymerization of Unicryl for the first time and
need advice about lamps, embedding molds which are UV penetrable, etc.,
etc. Virtually no specific advice is given in the circular I received or
in the two papers referenced in this literature. I would appreciate any
tips, and/or references from any experienced user. Thank you. Grace
Kennedy






From: mme-at-map.com (barbara foster)
Date: Mon, 31 Mar 1997 14:02:39 -0800
Subject: Re: Info on image analysis software/solutions for

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Luc Nocente wrote:
}
} ------------------------------------------------------------------------} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} -----------------------------------------------------------------------.}
} Try Visilog from Noesis Vision Inc, we have a turn key system available for
} sperm analysis and or fibre analysis. You can reach us at
} http://www.noesisvision.com
}
} At 12:51 PM 3/7/97 +0530, SONEJA A K wrote:
}
} ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
}
} -----------------------------------------------------------------------.
} }
} }
} } Hello guys,
} }
} } Looking for image analysis software/solutions for the following:
} }
} } 1} sperm analysis,cell motility etc.
} } 2} textile analysis,rayon fibre analysis,broken fibre etc.
} }
} } Could anyone point out some good manufacturers/sources with
} } address/email/fax wwith name of source.?//////
} } I would be grateful if someone could help me in this context.
} }
} } Thanks,
} }
} } Best regards,
} }
} } Anish
} }
}
} *************************************************************************
} } For further details please contact:
} } Soneja A.K.
} } Director
} } METZER BIOMEDICAL & ELECTRONICS PVT.LTD.
} } 327 Wadala Udyog Bhavan,Wadala,MUMBAI(BOMBAY )400 031.INDIA
} } Tel:91 22 4145057/4165650
} } Fax 91 22 4168757
} }
} } Email:soneja-at-giasbma.vsnl.net.in
}
} *************************************************************************
} }
} }
}
} ----------------------------------------------------------------------------
} ---------------------
} Luc Nocente Tel: 514 345 1400
} Noesis Vision Inc. Fax: 514 345 1575
} e-mail: ln-at-noesisvision.com
} 6800 Cote de Liesse, Suite 200
} St-Laurent, PQ
} H4T 2A7,Canada
}
} Visit our new web site at http://www.noesisvision.com
} ----------------------------------------------------------------------------
} ---------------------Dear Anish,

A belated response to your message of 3/7 re: image analysis for sperm
motility/analysis:

We did some work recently with Hamilton-Thorne Research. They have
several interesting, built-for-purpose systems. Contact:
Dr. Dairmid-Douglas Hamilton
100 Cummings Place, Suite 102C
181 Elliott Street
Beverly, MA 01915
Phone: (508)921-2050 Fax: (508)921-0250
Please feel free to mention my name.

Good luck.
Barbara Foster, Consortium President




From: leibest-at-acpub.duke.edu (Leslie Eibest)
Date: Mon, 31 Mar 1997 08:26:23 -0500
Subject: JEOL gold on carbon standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

John-
My JEOL parts catalog lists a gold on carbon standard. The part #
is 613149. Orders can be sent to JEOL (USA) Inc., Parts Dept., 11 Dearborn
Road, Peabody, MA 01960, or call (508) 535-5900 and ask for parts. I don't
have a recent price list, so I can't help you there.
Leslie

Leslie Eibest
Zoology Dept., Box 90325
Duke University
(919) 684-2547
leibest-at-duke.edu






From: Bob_Citron-at-cc.chiron.com (Bob Citron)
Date: 3/31/97 8:55 AM
Subject: Turbo Pumped SEM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

John:

We have used a turbo pumped SEM since 1986. Our experience with this system has
actually been quite good - we have only had to replace it once. It was
fortunate that we are under contract because the pump cost is rather high. The
greatest benefits that I see from this system are 1) pump-down time (usually
only a few minutes) and 2) cleanliness. One other thing to consider; on our
system, the turbo hangs down from a metal bellows connected to the base of the
chamber. At one time, we had a very small crack in this bellows, which was not
detected for over a year. We replaced numerous ion pumps and other components
until it was finally located. Again, fortunately we have always had the system
under contract.

-Bob
********************************
Bob Citron
Chiron Vision
555 W. Arrow Hwy
Claremont, CA 91711
PH: (909)399-1311
Email: Bob_Citron-at-cc.chiron.com
********************************

______________________________ Reply Separator _________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I would appriciate hearing from anyone on/off line with experience
with turbo pumped SEM's (benefits, horror stories, etc.)

Thank you

John Humenansky
Braun Interec Corp.
6875 Washington Ave. So.
Minneapolis, MN 55439
(612) 942-4822




From: jhumenansky-at-brauncorp.com
Date: 3/31/97 3:37 AM
Subject: SEM Jeol high res standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


=20


______________________________ Reply Separator ____________________________=
_____


------------------------------------------------------------------------=20
The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20
John Bozzola wrote:
=20
Jeol sells a high resolution standard (gold on carbon) that we would like=20
to purchase. Does anyone have info: (part number, price, where to order)?=20
Many thanks.
=20
####################################################################=20
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Neckers Building, Room 146 - B Wing
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html=20
####################################################################
=20
John;
You can contact JEOL parts directly in Peabody, MA or if you have the=20
right lab equipment you can make your own. A vacuum evaporator and a=20
30-40 amp. external power supply that can be connected to a second set=
=20
of evaporator electrodes is all the equipment that is required. The=20
supplies are standard EM supplies that can ordered from any of the=20
usual sources. If you are interested in making your own I'd be glad=20
to send more details on/off line.
=20
John Humenansky
Braun Intertec Corp.
6875 Washington Ave. So.
Minneapolis, MN 55439
(612) 942-4844
=20




From: heckman-at-pilot.msu.edu (John heckman)
Date: Mon, 31 Mar 1997 08:12:38 -0500
Subject: Re: SEM Jeol high res standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



John,

If you have a sputter coater or evaporator, why not make your own. I've
had good luck making calibration standards like this (I use them for camera
constant calculation, resolution, demos, in TEM). It's good practice in
thin film making for students, too ;-)

1. Make a thin carbon film by carbon coating a formvar, butvar etc. coated grid.
(how thin the carbon should be depends on the final use; if you want to see
the gold lattice in TEM it should be painfully thin). Remove the plastic
backing by placing it in a (glass) Petri dish lined with a filter paper pad
soaked in CHCl3 (hood).

2. After an overnight stint in the Petri dish, you should have mostly carbon
left on the grids.

3. Pop them in your sputter-coater and give them a short blast of Au (we've got
a pure Au target) if you use Au/Pd in yours, I'm not sure what you'll
get... Well, if you've got a gold target about 1/12 of your normal sputtering
time to coat an average SEM sample should work. You'll have a collection of
isolated islets of Au. Carbon tape it to a stub and have a look.

Hope this helps, even though it doesn't answer the question.

cheers,

John Heckman
TEM supervisor/Center for Electron Optics
Michigan State University

Disclaimer: The preceding technique works in my reality; follow my
instructions at your own risk!

} -----------------------------------------------------------------------.
}
} Jeol sells a high resolution standard (gold on carbon) that we would like
} to purchase. Does anyone have info: (part number, price, where to order)?
} Many thanks.
}
} ####################################################################
} John J. Bozzola, Ph.D., Director
} Center for Electron Microscopy
} Neckers Building, Room 146 - B Wing
} Southern Illinois University
} Carbondale, IL 62901-4402
} U.S.A.
} Phone: 618-453-3730
} Fax: 618-453-2665
} Email: bozzola-at-siu.edu
} Web: http://www.siu.edu/departments/shops/cem.html
} ####################################################################






From: Simon Watkins :      swatkins-at-pitt.edu
Date: Mon, 31 Mar 97 09:46:35 -0500
Subject: philips EM300

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

-- [ From: Simon Watkins * EMC.Ver #2.5.02 ] --

Hi folks
We have a em300 that is about to join the great trashcan in the sky. Before
it meets its demise I was wondering whether anyone wants parts from it,

let me know

simon

--
-------------------------------------------------
Simon C. Watkins Ph.D
Associate Professor
Director Center for Biologic Imaging
University of Pittsburgh
Pittsburgh PA 15261
tel 412-648-3051
fax 412-648-2004
-----------------------------------------------




From: jhumenansky-at-brauncorp.com
Date: Mon, 31 Mar 1997 8:55:15 -0600
Subject: Turbo Pumped SEM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

I would appriciate hearing from anyone on/off line with experience=20
with turbo pumped SEM's (benefits, horror stories, etc.)
=20
Thank you
=20
John Humenansky
Braun Interec Corp.
6875 Washington Ave. So.
Minneapolis, MN 55439
(612) 942-4822




From: bozzola-at-siu.edu (John J. Bozzola) (by way of Nestor J. Zaluzec)
Date: Sun, 30 Mar 1997 22:03:21 -0500
Subject: SEM Jeol high res standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jeol sells a high resolution standard (gold on carbon) that we would like
to purchase. Does anyone have info: (part number, price, where to order)?
Many thanks.

####################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Neckers Building, Room 146 - B Wing
Southern Illinois University
Carbondale, IL 62901-4402
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################


The JEOL resolution standard can be purchased from the JEOL USA parts
department. Phone # (508)535-5900. The part number is 613149. I
believe this is what you are asking for but verify that with them.

Disclaimer: I have no interest in JEOL

Michael D. Standing
e-mail: MDStandi-at-bioag.byu.edu
Phone: (801)378-4011




From: Beverly E Maleeff
Date: 31 Mar 97 18:18:50 EDT
Subject: April '97 PSM Meeting Notice

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

PHILADELPHIA SOCIETY FOR MICROSCOPY
APRIL 1997 MEETING NOTICE

DATE: Tuesday, April 15, 1997

PLACE: Laboratory for Research on the Structure of Matter
(LRSM) Building, 33rd and Walnut Street.
Parking is available behind the LRSM Building after 5:00 PM.

TIME: 5:30 PM Social hour, hosted by our meeting sponsor

6:30 PM Dinner (see menu at end of message)
Members $12.00 Student members $6.00 Non-members $15.00
RESERVATIONS REQUIRED!!!

7:30 PM Speaker

Chemistry in the Crime Lab
Richard Saferstein, Ph.D.

Forensic science in its broadest definition is the application of science to
law.
As our society has grown more complex, it has become more dependent on
rules of law to regulate the activities of its members. Forensic science offers
the knowledge and technology of science for the definition and enforcement
of such laws. The analytical techniques that can be used in forensic science
are numerous and diverse. In general, they must be sensitive enough to cope
with minute samples of physical evidence, but they must also be reliable and
reproducible to withstand scrutiny by fellow experts in and out of the
courtroom.
Speed and economy have to be considered too, for the typical forensic scientist
must analyze hundreds of cases each year.
In this talk a variety of analytical procedures that are applicable to solving
forensic
science problems will be presented. Included in the discussion will be the
application
of microscopic analysis to forensic hair and fiber. Also discussed will be the
role of
visible and infrared microspectrophotometric techniques in forensic problem
solving.
The speaker will review significant achievements that have been made in
utilizing
DNA typing for the purposes of linking blood and semen evidence to a single
individual.
A number of actual case discussions will be included in the talk in order to
exemplify
the relevancy of forensic science to criminal investigation.


Dinner Menu:
Beer, wine, assorted soda and bottled water
Munchies

Mini wontons with dipping sauce
Chicken stir fry
Vegetarian stir fry
Fried rice
Spinach salad

Cake du jour
Coffee, decaf or tea


NEWS FLASH: Reservations can now be sent by e-mail by writing to
PSM-RESERVATIONS-at-INAME.COM. Be sure to include your name
and affiliation in the body of the message. PSM prefers that you use this
convenient method to RSVP.

If you don't have access to e-mail, phone reservations will be taken by
Ms. Pat Overend at the University of Pennsylvania, 215/898-8337.
Deadline for reservations will be Friday, April 11. If you have any
questions regarding the meeting please feel free to contact Rollin Lakis
of the Executive Council. Cancellations must be received no later than
5:00 PM, April 14, 1997.
RESERVATIONS ARE REQUIRED FOR DINNER. We cannot
guarantee you a meal if you do not make a reservation by the deadline listed
above.










From: Melvyn Dickson :      M.Dickson-at-unsw.edu.au
Date: Tue, 01 Apr 1997 10:43:30 +1000
Subject: Re: Turbo Pumped SEM's

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

Message-Id: {1.5.4.32.19970401004330.00680654-at-pop3.unsw.edu.au}
X-Sender: s7001031-at-pop3.unsw.edu.au
X-Mailer: Windows Eudora Light Version 1.5.4 (32)
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"

We have two turbo pumped SEMs - a Leo/Cambridge S-360 (1988) and an Hitachi
S-900 (1991). We have never had a problem with either turbo. Local Leo
service had a policy of reconditioning the (Pfeiffer) turbos every 3 years
for bearing replacement just in case. Their policy now is to leave it until
there is a sign of imbalance (audible noise). They are confident now that
turbos are no more trouble than any other part so take no special
precautionary measures apart from re-oiling the bearing every 6 months. The
S-900 has a mag lev. Seiko Turbo. Hitachi advice is that so long as you
change the backup batteries every 12 months to ensure full charge (in case
of power failure the backup keeps the turbine levitated until it runs down)
- the pump will last forever.


Advantage is no fuss clean vacuum. The source of contamination has to be
your specimen (or some backstreaming from the rotary, though oils oughtn't
get through the turbo if its runnning). AND no problem with water
supply/cooling and very little problem with power failure. (Everything
coasts to a stop).

BUT a well designed oil diffusion pump system can be NEARLY as good. For
that reason I did not insist on the turbo option with our latest Hitachi
S-4500. How sure are you of design? Some well known SEMs were plagued with a
badly designed diff pump system which stalled and backstreamed with nearly
every specimen change!

Mel Dickson






From: Rafal Spirydon :      spirydon-at-matlb.kjist.ac.kr
Date: Tue, 1 Apr 1997 10:33:01 +0900
Subject: Debye-Waller temperature factor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am interested in HREM image simulations of III-V semiconductors and I am
looking for any references where I can find the values of the Debye-Waller
temperature factors for such atoms as P, Ga, IN, As etc.
I would be extremely grateful for any helpful information

Sincerely yours

Rafal Spirydon