From: Tiina_Hallamaa-at-kcl.fi (Tiina Hallamaa)
Date: Tue, 01 Dec 1998 08:47:07 +0200
Subject: constant brightness/contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,

We are looking for a way to set the contrast and the brightness values in
the SEM's backscatter mode at the same level from day to day and rather so
that any user can set it to this wanted level. We have not found a good
way of doing this. Any suggestions?

Thanks a lot,


Tiina

*************************************************************
Tiina Hallamaa, M.Sc (Eng)
The Finnish Pulp and Paper Research Institute
P.O.Box 70
FIN-02151 Espoo
Finland
*************************************************************
phone: +358-9-4371 492, fax: +358-9-464 305
email: Tiina.Hallamaa-at-kcl.fi









From: Giles, John E Jr (FL51) :      jegiles-at-space.honeywell.com
Date: Tue, 1 Dec 1998 05:01:32 -0500
Subject: Re: Constant brightness/contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tiina,

We use a waveform monitor on our AMRAY 1830. This allows us to set a pretty
constant brightness and contrast level so that photographs will come out
with the proper levels. This is much better than the auto brightness and
contrast or small row of LED's that our SEM had as a factory meter. The
waveform monitor resembles a small oscilloscope and has a fairly fine
pitched grid on the screen. Your SEM manufacturer should be able to supply
you with one. It has worked very well for us.

John Giles
Principal Materials Engineer
Honeywell Space Systems

} Dear all,

} We are looking for a way to set the contrast and the } brightness values in
the SEM's backscatter mode at the } same level from day to day and rather so
that any user } can set it to this wanted level. We have not found a } good
way of doing this. Any suggestions?

} Thanks a lot,

} Tiina






From: jiahui yu :      ydyu-at-eudoramail.com
Date: Tue, 01 Dec 1998 04:47:53 -0700
Subject: How to find the good orientation to recored Ceramic GB HRTEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ceramic TEM experts and folks,


Now I want to record the HRTEM images of Ceramic Grain Boundaries, but I found it is very difficult to find out a suitable sample area for both grains in the suitable orientations. Could you give me some advices about how to find them?

Many thanks,

---
Best Wishes,

JH Yu


Join 18 million Eudora users by signing up for a free Eudora Web-Mail account at http://www.eudoramail.com





From: Woody.N.White-at-mcdermott.com
Date: Tue, 1 Dec 1998 7:56:00 -0600
Subject: Re: constant brightness/contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html







Tiina,

If your SEM has a video waveform display available, it can be used
with standards to consistently set BSE contrast (gain) and
brightness (dc level). The best standards should be polished flat
and the atomic number should bracket the anticipated BSE range
required. Use the standards to set desired black/white levels.

Woody White
McDermott Technology, Inc.


Dear all,

We are looking for a way to set the contrast and the brightness values in
the SEM's backscatter mode at the same level from day to day and rather so
that any user can set it to this wanted level. We have not found a good
way of doing this. Any suggestions?

Thanks a lot,


Tiina

*************************************************************
Tiina Hallamaa, M.Sc (Eng)
The Finnish Pulp and Paper Research Institute
P.O.Box 70
FIN-02151 Espoo
Finland
*************************************************************
phone: +358-9-4371 492, fax: +358-9-464 305
email: Tiina.Hallamaa-at-kcl.fi





From: jiahui yu :      ydyu-at-eudoramail.com
Date: Tue, 01 Dec 1998 04:48:18 -0700
Subject: How to find the good orientation to recored Ceramic GB HRTEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Ceramic TEM experts and folks,


Now I want to record the HRTEM images of Ceramic Grain Boundaries, but I found it is very difficult to find out a suitable sample area for both grains in the suitable orientations. Could you give me some advices about how to find them?

Many thanks,

---
Best Wishes,

JH Yu


Join 18 million Eudora users by signing up for a free Eudora Web-Mail account at http://www.eudoramail.com





From: jrnelson :      jrnelson-at-nj1.aae.com
Date: Tue, 01 Dec 1998 09:16:00 -0500
Subject: Re: background x-ray radiation and maximum permissible dose of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


First you should obtain NJDEP's Chapter 28 covering Radiation Protection Programs
to see what rules need to be followed. New Jersey is very strict about this topic.
The current edition 5/15/95 is good for five years.

A summary table in "7:28-6.1 Permissible dose rates, radiation levels and
concentrations" indicates 1.25 rems per year. However, this section is long with
many exceptions. If you send me an E-Mail with the details of your question, I
will do my best to answer it.

J. Roy Nelson
Material Testing Laboratory
Pennington, NJ
(609) 730-0575
jrnelson-at-nj1.aae.com

narahari ramanuja che stnt wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hello,
} we are doing some research on protective shields for x-ray
} radiation. does anyone know what is the permissible or background
} radiation intensity? also on refering to an old publication by the British
} Journal of Medicine, i found out that the maximum permissible dose(MPD)
} for exposure of the whole body to radiation expressed in rontgens is 0.3
} per week, has this level changed or is it still the same? to calculate the
} MPD, we need to know the dose rate (in r min), does anyone have some info
} on this?
}
} it would be great if you could give me some info on these topics.
}
} sincerely,
}
} Narahari.
}
} ________________________________________________________________________________
}
} Narahari Ramanuja
}
} Office : Materials Synthesis & Characterization Lab,
} Dept of Materials Science & Engineering,
} NJIT,
} ph # 201-596-3680
} email : nxr0308-at-megahertz.njit.edu
} ________________________________________________________________________________
}






From: Jim J Darley :      jim-at-proscitech.com.au
Date: Wed, 2 Dec 1998 00:11:34 +1100
Subject: RE: constant brightness/contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tiina:
SEM's have a linescan mode and this can be centred and
modulated. The amplitude in the modulation then represents
contrast and the position of the graph (which is centred
before modulation) indicates brightness. This is easy
enough to calibrate and teach to users. Modern instruments
have various aids to obtain good exposures, some are
simpler but they are never better than that modulation.
Just fixing the brightness and contrast controls is rather
more complicated, because this would necessitate fixing all
working parameters in an SEM, always the same: specimens
and coatings, WD, condenser setting and apertures, emission
and specimen tilt. The list is endless, even the
scintillator cannot age. Forget fixing just brightness and
contrast amplifications and try to give users at least
minimal instructions in SEM operations and that should
include the modulated linescan.
Cheers
Jim Darley
ProSciTech Microscopy
PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 7 4774 0370 Fax: +61 7 4789 2313
Great microscopy catalogue, 500 Links, MSDS, User Notes
********************** www.proscitech.com.au *****



On Tuesday, December 01, 1998 4:47 PM, Tiina Hallamaa
[SMTP:Tiina_Hallamaa-at-kcl.fi] wrote:
}
----------------------------------------------------------
} --------------
} The Microscopy ListServer -- Sponsor: The Microscopy
} Society of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.ht
} ml
}
----------------------------------------------------------
} -------------.
}
}
} Dear all,
}
} We are looking for a way to set the contrast and the
} brightness values in
} the SEM's backscatter mode at the same level from day to
} day and rather so
} that any user can set it to this wanted level. We have
not
} found a good
} way of doing this. Any suggestions?
}
} Thanks a lot,
}
}
} Tiina
}
}
**********************************************************
} ***
} Tiina Hallamaa, M.Sc (Eng)
} The Finnish Pulp and Paper Research Institute
} P.O.Box 70
} FIN-02151 Espoo
} Finland
}
**********************************************************
} ***
} phone: +358-9-4371 492, fax: +358-9-464 305
} email: Tiina.Hallamaa-at-kcl.fi
}
}
}
}






From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Tue, 01 Dec 1998 08:42:06 -0600
Subject: Re: constant brightness/contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Just to expand on Woody's point, we use some TV tube coat (C) sprayed on a
chip of calcite as a thin, portable standard to set the BSE brightness
levels for concrete analyses. We simply lay the standard on the sample as
we insert it into our microscope. Of course, one could also embed samples
of brass or calcite or silica in epoxy and polish them flat to do the same
thing. You could then have multiple reference levels.

Of course, some sort of waveform display is required. We have ours marked
so that we can set the levels reproducibly.

Warren

At 07:56 AM 12/1/98 -0600, Woody wrote:
}
} Tiina,
}
} If your SEM has a video waveform display available, it can be used
} with standards to consistently set BSE contrast (gain) and
} brightness (dc level). The best standards should be polished flat
} and the atomic number should bracket the anticipated BSE range
} required. Use the standards to set desired black/white levels.
}
} Woody White
} McDermott Technology, Inc.
}
} Dear all,
}
} We are looking for a way to set the contrast and the brightness values in
} the SEM's backscatter mode at the same level from day to day and rather so
} that any user can set it to this wanted level. We have not found a good
} way of doing this. Any suggestions?
}
} Thanks a lot,
}
} Tiina







From: drose-at-wlgore.com
Date: Tue, 1 Dec 1998 11:34:22 -0500
Subject: Profilometry - Standards available?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear List,

A colleague is interested in a standard to evaluate a variety of
profilometry instruments. Both metallic and polymeric standards would be
of use. X, Y, and Z directions would need to be measured, something in the
100-1000um range. Also, does anyone know a manufacturer who will make
polystyrene latex spheres at a requested diameter?

TIA

=======================
David Rose
WL Gore and Associates
297 Blue Ball Road
Elkton, MD 21921
=======================







From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Tue, 01 Dec 1998 08:49:27 -0800
Subject: Re: background x-ray radiation and maximum permissible dose of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Narahari,
The usual upper limit that is applied for any radiation-emitting devices in
Canada and the USA is "0.5 mR/hr. at a distance of 5 cm from any part of
the device". This is from the "Electron Microscopy Safety handbook" second
edition, Barber and Mascorro editors. This is an excellent, slim, soft-cover
book that should be in any EM lab, IMHO. There is a lot of other info on
background radiation and other radiation hazards in the EM lab.
You wrote:
} Hello,
} we are doing some research on protective shields for x-ray
} radiation. does anyone know what is the permissible or background
} radiation intensity? also on refering to an old publication by the British
} Journal of Medicine, i found out that the maximum permissible dose(MPD)
} for exposure of the whole body to radiation expressed in rontgens is 0.3
} per week, has this level changed or is it still the same? to calculate the
} MPD, we need to know the dose rate (in r min), does anyone have some info
} on this?
}
} it would be great if you could give me some info on these topics.
}
} sincerely,
}
} Narahari.
}
} ___________________________________________________________________________
_____
}
} Narahari Ramanuja
}
} Office : Materials Synthesis & Characterization Lab,
} Dept of Materials Science & Engineering,
} NJIT,
} ph # 201-596-3680
} email : nxr0308-at-megahertz.njit.edu
} ___________________________________________________________________________
_____
}
}
}
}
}
}
}
}
}
}
}
}
}
}
}
}
}
Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
fax: 604-822-3619
e-mail: mager-at-interchg.ubc.ca






From: Marti, Jordi :      Jordi.Marti-at-alliedsignal.com
Date: Tuesday, December 01, 1998 6:48AM
Subject: How to find the good orientation to recored Ceramic GB HRTEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


JH YU:

I spent some time doing the same type of work in beta silicon nitride , and
yesI found it very time consuming because one needs to find orientations
such that the two grains are in orientations that can be imaged and at
the same time the grain boundary needs to be parallel to the beam. You might
find the paper by D.R. Clark (Ultramicroscopy 4 (1979)33-44) useful.

What I found to be helpful was to first check the image of the boundary at
lower mags (typically X100K) in BF mode and tilt about an axis parallel to
the image of the boundary until I attained a relatively sharp "minimum"
in the width of the boundary .This is relatively fast and it allows you to
eliminate those GBs that are way out of orientation ( I was using a 30
degree double tilt holder).

Once I had what seemed to be a suitable GB, then,at that point I would
check the diffraction pattern to see if there were any Kikuchi lines (from
either grain) that were parallel to the grain boundary and which
corresponded to reflections that could be used for HRTEM imaging (i.e. large
enough d-spacing). If that was the case, then I would tilt the sample along
that Kikuchi line (i.e. maintaining that set of reflections for that
particular grain) and at the same time check the DP from the other grain
until a suitable orientation was attained for that second grain. More often
than not, I had to move on to another location and start the whole
operation again until all three elements(Grain 1, grain 2 and the GB) had
the proper orientation.

In most instances my images consisted of two beam conditions but that was
sufficient to resolve the GBs which were typically about 1 to 4 nm wide.
In one afternoon session I could probably end up with three to five
suitable boundaries (plus coffee breaks).

I hope this helps.

Jordi Marti


----------
} From: jiahui yu
To: For sending TEM discussion
-----------------------------------------------------------------------.


Dear Ceramic TEM experts and folks,


Now I want to record the HRTEM images of Ceramic Grain Boundaries, but I
found it is very difficult to find out a suitable sample area for both
grains in the suitable orientations. Could you give me some advices about
how to find them?

Many thanks,

---
Best Wishes,

JH Yu


Join 18 million Eudora users by signing up for a free Eudora Web-Mail
account at http://www.eudoramail.com






From: William Tivol :      tivol-at-wadsworth.org
Date: Tue, 01 Dec 1998 13:15:59 -0500
Subject: Re: background x-ray radiation and maximum permissible dose of radiation.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


narahari ramanuja che stnt wrote:
}
} we are doing some research on protective shields for x-ray
} radiation. does anyone know what is the permissible or background
} radiation intensity? also on refering to an old publication by the British
} Journal of Medicine, i found out that the maximum permissible dose(MPD)
} for exposure of the whole body to radiation expressed in rontgens is 0.3
} per week, has this level changed or is it still the same? to calculate the
} MPD, we need to know the dose rate (in r min), does anyone have some info
} on this?
}
} it would be great if you could give me some info on these topics.
}
Dear Narahari,
That must be a very old book. The current standards (at least
in New York State) are 0.5 rem/year (about equal to 0.5 roentgens/year)
for the general public and 5 rem/year for radiation workers; however,
for women who are or may be pregnant, the lower level applies even for
radiation workers. The dose rate must be measured with an ionization
chamber detector (we use a Model 2588 from Nuclear Chicago). Due to
the unevenness of shielding by the radiation source itself (e.g., the
EM column), a calculation will not usually be useful. You should
also operate the equipment to produce the worst possible radiation
field when making these measurements--put the EM beam onto as many
high-Z parts as possible, etc.--so that even operator mistakes will
not result in exceeding the permissible levels. Good luck.
Yours,
Bill Tivol





From: Lehman, Ann :      Ann.Lehman-at-exchange.cc.trincoll.edu
Date: Tue, 1 Dec 1998 14:08:12 -0500
Subject: Philips 6585 STEM Search

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,

A Philips EM400T w/ 6585 STEM was donated to our new EM teaching facility
last year, to establish TEM & analytical capabilities
(biology/engineering/physics/chemistry). On PEI's recommendation, we
invested around $30K to move & refurbish the instrument, plus more to
upgrade the EDS system.

End result:
The column was in good shape to begin with, and now it is great (GO,
Philips). However, after expending the greater part of the available funds
in man-hours and parts on the STEM unit over a period of months, the STEM
was merely transformed into a(n expensive!) heap of garbage that nothing can
resuscitate (BOO, Philips).

Does anyone out in ListServerLand have a 6585 STEM unit you might loan or
donate to our program?

PEI has offered to transport & install it for us (GO, Philips). Having put
so much into our failed STEM, and remaining committed to the concept, we can
also spend a little more for a(n operating!) unit.

Thanks, all.

Ann Hein Lehman
TEM Lab Mgr, Trinity College
LSC 314
300 Summit St.
Hartford, CT 06106
860-297-4289
Ann.Lehman-at-exchange.cc.trincoll.edu





From: Heike Grabsch :      h.grabsch-at-uni-koeln.de
Date: Tue, 01 Dec 1998 20:48:04 -0800
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: Marty Reed :      mmr7001-at-axe.humboldt.edu
Date: Tue, 01 Dec 1998 12:11:33 -0800
Subject: sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Our sputter coater (ISI PS-2) just died. I need parts for it and was
hoping that someone may have an old one setting around that I could get for
parts. I would also consider a used sputter coateer of any make that you
may want to sell.

Thank you

Marty Reed
Equipment Technician
Biology Department
Humboldt State University
Arcata CA 95521
707-826-3234
707-826-3201 FAX
mmr7001-at-axe.humboldt.edu





From: Michael Bode :      mb-at-soft-imaging.com
Date: Tue, 1 Dec 1998 14:04:49 -0700
Subject: RE: constant brightness/contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tiina,
}
} There are a number of things you can do, depending on whether you want to
}
} a) just acquire images with the same contrast and brightness to facilitate
} comparison or
} b) want to acquire BSE images and get materials data from these. b) would
} have to be split up further into
}
} i) image comparison
} ii) semi-quantitative data without standard
} iii) semi-quantitative with standard
}
} a) if you just want to get images with similar brightness and contrast, the
} easiest and least expensive way would be to use the y-modulation or linescan
} mode of the SEM, adjust brightness and contrast so that the amplitudes are
} similar for different images, and take an image.
}
} b, i) if you want to use the Z-contrast of the SEM to obtain information
} about material distribution in your sample, you probably have to pay more
} attention to other parameters such as WD, beam current, contamination, etc.
} Again, if you are just looking for a comparison of two areas you may be able
} to use the method a).
}
} b,ii) Now it get's more interesting. If you want to obtain semi-quantitative
} data you must quantify the gray levels of the image, which can then in turn
} in principle be used to measure Z. If you don't want to use a standard, you
} probably have to fix most of your parameters (WD, beam current, tilt angles,
} etc). Once that is done, you need to acquire the images digitally, at which
} point you can calibrate the intensity and then have the software threshold
} different intensity levels and match them to different materials. The
} matching could be done either by using theoretical values of I(z), or by
} using values determined from known materials.
}
} b,iii) If you can use a standard in the SEM at the same time as the sample,
} this would give you probably the best results. Acquire an image of the
} standard, then of the sample under the same conditions. Software can then be
} used to match the intensity of the standard to intensity on the sample and
} obtain information about the material. Again, you could then use software to
} threshold the intensity and tell you which material is where.
}
} I hope, that helped. If you have further questions regarding the digital
} processing part, contact me through email. We have people in Finland who can
} help you further.
}
} Michael Bode
}
} *******************************************************
} Disclaimer:
} Soft Imaging System produces and sells image acquisition
} and processing systems. We therefore have a vested
} interest in some of the items mentioned above.
} *******************************************************
}
}


Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
phone: (888) FIND SIS
fax: (303) 234-9271
email: info-at-soft-imaging.com

}
}
} On Tuesday, December 01, 1998 4:47 PM, Tiina Hallamaa
} [SMTP:Tiina_Hallamaa-at-kcl.fi] wrote:
} }
} ----------------------------------------------------------
} } --------------
} } The Microscopy ListServer -- Sponsor: The Microscopy
} } Society of America
} } To Subscribe/Unsubscribe -- Send Email to
} } ListServer-at-MSA.Microscopy.Com
} } On-Line Help
} } http://www.msa.microscopy.com/MicroscopyListserver/FAQ.ht
} } ml
} }
} ----------------------------------------------------------
} } -------------.
} }
} }
} } Dear all,
} }
} } We are looking for a way to set the contrast and the
} } brightness values in
} } the SEM's backscatter mode at the same level from day to
} } day and rather so
} } that any user can set it to this wanted level. We have
} not
} } found a good
} } way of doing this. Any suggestions?
} }
} } Thanks a lot,
} }
} }
} } Tiina
} }
} }
} **********************************************************
} } ***
} } Tiina Hallamaa, M.Sc (Eng)
} } The Finnish Pulp and Paper Research Institute
} } P.O.Box 70
} } FIN-02151 Espoo
} } Finland
} }
} **********************************************************
} } ***
} } phone: +358-9-4371 492, fax: +358-9-464 305
} } email: Tiina.Hallamaa-at-kcl.fi
} }
} }
} }
} }
}
}
}





From: Randy Tindall :      rtindell-at-NMSU.Edu
Date: Tue, 01 Dec 1998 15:22:01 -0700
Subject: lubricant for cryostat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Does anyone have recommendations for a low-temperature lubricant for our
old IEC cryostat. The literature that we have available just refers to
"special low-temperature oil, product no. ####". We are thinking of
molybdenum sulfide or graphite, but if there is any really good recommended
lubricant for -30 degrees C, that's what we want.

Thanks in advance.

Randy


Randy Tindall
Electron Microscope Laboratory
Box 3EML
New Mexico State University
Las Cruces, NM 88003

rtindell-at-nmsu.edu (work)
nrtindall-at-zianet.com (home)





From: Terry Cooper :      terry.cooper-at-btinternet.com
Date: Wed, 2 Dec 1998 09:55:26 -0000
Subject: TAAB Address etc

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi People,

Having been off line for a while I see we have managed to confuse most of
the world
thanks to British Telecom. For Scott who asked about flat bottom vials we
do
manufacture these and much else besides. Stephen Griffiths from the
Institute of Ophthalmology got it right although sadly there are no prizes
for him this time. Anyone requiring details can contact us as follows:

TAAB Laboratories Equipment Ltd, 3 Minerva House, Calleva Park, Aldermaston
Berkshire, RG7 8NA, England
Tel ++44 (0) 118 981 7881
Fax 110 981 7881
Company e-mail sales-at-taab.co.uk

Regards to all

Terry Cooper





From: Jim J Darley :      jim-at-proscitech.com.au
Date: Wed, 2 Dec 1998 13:55:21 +1100
Subject: RE: lubricant for cryostat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Randy: I agree that moly sulfide or graphite would be
satisfactory.
However, Apiezon N grease is designed for medium and very
low temperature applications. It's sticky and a very thin
film would stay in place for a long time.
You can read about Apiezon N properties in our online by
going to page M2 from the contents page.
Disclaimer: ProSciTech is a supplier and sells this
products.
Cheers
Jim Darley
ProSciTech Microscopy
PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 7 4774 0370 Fax: +61 7 4789 2313
Great microscopy catalogue, 500 Links, MSDS, User Notes
********************** www.proscitech.com.au *****

}
} Hi,
}
} Does anyone have recommendations for a low-temperature
} lubricant for our
} old IEC cryostat. The literature that we have available
} just refers to
} "special low-temperature oil, product no. ####". We are
} thinking of
} molybdenum sulfide or graphite, but if there is any
really
} good recommended
} lubricant for -30 degrees C, that's what we want.
}
} Thanks in advance.
}
} Randy
}
}
} Randy Tindall
} Electron Microscope Laboratory
} Box 3EML
} New Mexico State University
} Las Cruces, NM 88003
}
} rtindell-at-nmsu.edu (work)
} nrtindall-at-zianet.com (home)






From: A G DU TOIT :      AGDT-at-GERGA.SUN.AC.ZA
Date: Wed, 2 Dec 1998 07:28:29 -0500
Subject: Re: lubricant for cryostat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
We got ours from LIPSHAW and it is called
LIPSHAW CRYOTOME LUBRICANT No. 291.
The address
Lipshaw Corp.
7446 Central Avenue
Detroit. MI 48210.
Hope this will help.
Regards.
Andre du Toit
Department of Anatomical Pathology
Tygerberg Hospital
Cape Town
South Africa.







From: Terry Cooper :      terry.cooper-at-btinternet.com
Date: Wed, 2 Dec 1998 17:14:03 -0000
Subject: TAAB Address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,

We seem to have confused the entire world thanks to British Telecom.

We can supply flat bottom capsules as requested by Scott and other things
too.

We can be contacted as below:

TAAB Laboratories Equipment Ltd
3 Minerva House, Calleva Park
Aldermaston, Berkshire RG7 8NA, England

Tel ++44 (0) 118 981 7775
Fax 118 981 7881
Company e-mail sales-at-taab.co.uk

Regards

Terry Cooper





From: Caroline Schooley :      schooley-at-mcn.org
Date: Wed, 2 Dec 1998 09:27:38 -0800 (PST)
Subject: Food microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A new web page was announced recently on this listserver; it has
informative images and text that will help those who are doing educational
outreach. It has just been expanded to include some difficult-to-access
information:

The FOODS UNDER THE MICROSCOPE Web page (http://www.cyberus.ca/~scimat/)
has been extended and now contains the TABLE OF CONTENTS of the FOOD
STRUCTURE (formerly FOOD MICROSTRUCTURE) journal published in 1982-93. This
particular page may be accessed separately at
http://www.cyberus.ca/~scimat/journal.htm

Please send inquiries directly to Milos; he isn't a list subscriber.

Caroline Schooley
Educational Outreach Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.MSA.microscopy.com/ProjectMICRO/Books.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html







From: Edoardo Bemporad :      e.bemporad-at-materials10.dimi.uniroma3.it
Date: Wed, 2 Dec 1998 15:56:00 +0100
Subject: NiTi etchant

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This message is in MIME format. Since your mail reader does not understand
this format, some or all of this message may not be legible.

------ =_NextPart_001_01BE1E1D.87310D90
Content-Type: text/plain

Any suggestion on proper etchant procedure to use in order to do NiTi
Memory Shape alloys SEM images?
Thank You in advance, Edoardo.

------ =_NextPart_001_01BE1E1D.87310D90
Content-Type: text/html

{!DOCTYPE HTML PUBLIC "-//W3C//DTD HTML 3.2//EN"}
{HTML}
{HEAD}
{META HTTP-EQUIV="Content-Type" CONTENT="text/html; charset=us-ascii"}
{META NAME="Generator" CONTENT="MS Exchange Server version 5.5.1960.3"}
{TITLE} NiTi etchant {/TITLE}
{/HEAD}
{BODY}

{P} {FONT SIZE=2 FACE="Arial"} Any suggestion on proper etchant procedure to use in order to do NiTi Memory Shape alloys SEM images? {/FONT}
{BR} {FONT SIZE=2 FACE="Arial"} Thank You in advance, Edoardo. {/FONT}
{/P}

{/BODY}
{/HTML}
------ =_NextPart_001_01BE1E1D.87310D90--





From: Lesley Weston :      lesley-at-interchange.ubc.ca
Date: Wed, 2 Dec 1998 10:26:29 -0800 (PST)
Subject: Parts for Sorvall microtome

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anybody know where I can get parts for a Sorvall Porter-Blum MT II
ultramicrotome? If possible somewhere here in Canada, because of the exchange
rate. I think there have been questions like this in the past, but I can't seem
to access the archives. Thanks.

Lesley Weston.








From: Walck. Scott D. :      walck-at-ppg.com
Date: Tuesday, December 01, 1998 5:22PM
Subject: lubricant for cryostat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm not sure if graphite will work at those temperatures. Graphite gets
its lubrication properties from the layers being intercalated with gas
molecules. That is why it doesn't work in vacuum. MoS2 or WS2 works well
at low temperatures for tribological applications. It also works well in
vacuum applications. You probably will have to burnish it on and that could
create some debris and there is debris that can occur during usage.
-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Guys Run Rd. (packages)
P.O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)


"The opinions expressed are those of Scott D. Walck and not of PPG
Industries, Inc. nor of any PPG-associated companies."

----------
} From: Randy Tindall
To: microscopy-at-Sparc5.Microscopy.Com
-----------------------------------------------------------------------.


Hi,

Does anyone have recommendations for a low-temperature lubricant for our
old IEC cryostat. The literature that we have available just refers to
"special low-temperature oil, product no. ####". We are thinking of
molybdenum sulfide or graphite, but if there is any really good recommended
lubricant for -30 degrees C, that's what we want.

Thanks in advance.

Randy


Randy Tindall
Electron Microscope Laboratory
Box 3EML
New Mexico State University
Las Cruces, NM 88003

rtindell-at-nmsu.edu (work)
nrtindall-at-zianet.com (home)





From: kennedy-at-nsi.edu (Grace Kennedy)
Date: Wed, 2 Dec 1998 12:03:05 -0800
Subject: cryostat lubricant

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Proper low-temperature cryostat oil is available from the major suppliers
(i.e.Fisher), and is probably the best thing to use. My own experience is
that other lubricants usually become impossibly thick at low temperatures.








From: South Bay Technology :      Henriks-at-CompuServe.COM
Date: Wed, 2 Dec 1998 15:02:52 -0500
Subject: NiTi etchant

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Edoardo:

I do not have the answer for you, but there is a web site that has a
database of etchants that may be of help.

You can find it at:

www.kaker.com/mvd/vendors.html

I hope this helps!

Best regards-

David =

Writing at 11:34:28 AM on 12/2/98
=

*************************************************************************=
**
************************

David Henriks TEL: =

800-728-2233 (toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499=

San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com=


*************************************************************************=
**
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.


Message text written by Edoardo Bemporad
}
Any suggestion on proper etchant procedure to use in order to do NiTi
Memory Shape alloys SEM images?
Thank You in advance, Edoardo.
{





From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Wed, 2 Dec 1998 14:28:03 -0600 (CST)
Subject: attention venders

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




I am looking to purchase several items. A cryostat, a paraffin prosessor,
a freeze-substitution unit, and a freeze dryer. We need state-of-the-art
equipment. Please contact me at one of the addresses below.

Thank you.

Kathy Walters / /
Research Assistant III / /\
Center for Microscopy Research / /\ \
University of Iowa /_/ \ \
85 EMRB ____ ((O))
Iowa City, Iowa 52242 | | / /
|| / /
email: kwalters-at-emiris.iaf.uiowa.edu -----------
fax: (319)335-8049 -------------
www: http://www.uiowa.edu/~cemrf

Talk to me about joining the Iowa Microscopy Society.






From: Sara Miller :      saram-at-acpub.duke.edu
Date: Wed, 2 Dec 1998 15:41:05 -0500 (EST)
Subject: used chiller

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a TEM chiller (s)he wants to get rid of?

Thanx,
Sara


Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735






From: John Shields :      jpshield-at-arches.uga.edu
Date: Wed, 2 Dec 1998 16:54:53 -0500 (Eastern Standard Time)
Subject: Need info on GFP filters

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please respond to me directly, rather than on the listserve.
Thank you -
jpshield-at-arches.uga.edu

We are in the process of purchasing GFP filter(s) for Nikons (TE300
and Optiphot) we have at the Center. This is a multiuser facility,
so naturally we sould like to have a good bit of versatility in the
filter cube (if possible).
Any suggestions or comments regarding wavelengths, etc... would be
appreciated.
Thanks
********************************************
John P. Shields
Center for Ultrastructural Research
Barrow Hall
University of Georgia
Athens, GA 30602
(706)542-4080
jpshield-at-arches.uga.edu
********************************************






From: Kathy Walters :      kwalters-at-emiris.iaf.uiowa.edu
Date: Wed, 2 Dec 1998 16:36:34 -0600 (CST)
Subject: EELS for hire?????

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi, I'm looking for someone who might be interested in doing some work for
me. I am looking for sulpher in some cell membranes. You may contact me
at the following address if interested.

Kathy Walters / /
Research Assistant III / /\
Center for Microscopy Research / /\ \
University of Iowa /_/ \ \
85 EMRB ____ ((O))
Iowa City, Iowa 52242 | | / /
|| / /
email: kwalters-at-emiris.iaf.uiowa.edu -----------
fax: (319)335-8049 -------------
www: http://www.uiowa.edu/~cemrf







From: Tom Phillips :      tphillips-at-biosci.mbp.missouri.edu
Date: Wed, 2 Dec 1998 17:26:17 -0600
Subject: For Fuji Pictrography users only

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Our beloved Fuji Pictrography 3000 has developed a minor glitch. At a
reproducible spot in the image, there is a white (or almost white ) line
traveling from top to bottom (in the direction of the short width on the 8
x 5" prints). You can only see it in the gray areas of the print and it
doesn't appear to traverse solid black areas (we have moved a black bar
into the middle of the spot and don't see it passing thru but it is on
either side). This is not a straight line but looks like a squiggly line
with an amplitude of a couple of millimeters (as if a pen vibrated across
the surface as the paper was traveling under it). Has anybody experienced
this problem? More importantly, has anybody solved it? Thanks in advance.
Tom

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility

3 Tucker Hall
Division of Biological Sciences
University of Missouri
Columbia, MO 65211
(573)-882-4712 (voice)
(573)-882-0123 (fax)





From: Sally Stowe :      STOWE-at-rsbs.anu.edu.au
Date: Thu, 3 Dec 1998 17:46:00 +1000 GMT
Subject: wanted: Philips 430 film holders

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
Does anyone have spare cut film holders (36 plate) for Philips 400
series TEMs?

thanks
Sally


----------------------------------------------------------------------
Dr Sally Stowe |Email: stowe-at-rsbs.anu.edu.au
Facility Coordinator |Post: GPO Box 475
ANU Electron Microscopy Unit |ANUEMU (RSBS)
Ph 61 (0)2 6249 2743 |Australian National Univ.
FAX 61 (0)2 6249 4891 |Canberra, Australia 2601
http://online.anu.edu.au/EMU/home.htm







From: rschoonh-at-sph.unc.edu
Date: Thu, 03 Dec 1998 07:44:20 -0500 (Eastern Standard Time)
Subject: Re: RE: lubricant for cryostat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Cryostat lubricants can be found in the Fisher and VWR catalogs in the
Histology sections

regards,
Bob
Robert Schoonhoven
Laboratory of Molecular Carcinogenesis and Mutagenesis
Dept. of Environmental Sciences and Engineering
University of North Carolina
CB#7400
Chapel Hill, NC 27599
Phone
office 919-966-6343
Lab 919-966-6140
Fax 919-966-6123

**I'm willing to make the mistakes if someone else is willing to learn
from them**






From: Russell E. Cook :      cook-at-aaem.amc.anl.gov
Date: Thu, 3 Dec 1998 09:29:25 -0600
Subject: Centaurus BSD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Who manufacturesand/or markets the Centaurus BSE detector?

Russell E. Cook
Electron Microscopy Center for Materials Research
Argonne National Laboratory
Building 212
9700 South Cass Avenue
Argonne, IL 60439
(630)252-7194
FAX: (630)252-4798







From: Tong Wang :      tong-at-jlab.org
Date: Thu, 3 Dec 1998 11:50:07 -0500
Subject: fine metal tip used in microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,
Anybody knows where to buy some metal tips of 10 micron and 1 micron
diameter? Thanks in advance.

Tong







From: jrnelson :      jrnelson-at-nj1.aae.com
Date: Thu, 03 Dec 1998 12:21:28 -0500
Subject: Re: background x-ray radiation and maximum permissible dose of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


First you should obtain NJDEP's Chapter 28 covering Radiation Protection Programs
to see what rules need to be followed. New Jersey is very strict about
this topic.
The current edition 5/15/95 is good for five years.

A summary table in "7:28-6.1 Permissible dose rates, radiation levels and
concentrations" indicates 1.25 rems per year. However, this section is
long with
many exceptions. If you send me an E-Mail with the details of your
question, I
will do my best to answer it.

J. Roy Nelson
Material Testing Laboratory
Pennington, NJ
(609) 730-0575
jrnelson-at-nj1.aae.com

narahari ramanuja che stnt wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hello,
} we are doing some research on protective shields for x-ray
} radiation. does anyone know what is the permissible or background
} radiation intensity? also on refering to an old publication by the British
} Journal of Medicine, i found out that the maximum permissible dose(MPD)
} for exposure of the whole body to radiation expressed in rontgens is 0.3
} per week, has this level changed or is it still the same? to calculate the
} MPD, we need to know the dose rate (in r min), does anyone have some info
} on this?
}
} it would be great if you could give me some info on these topics.
}
} sincerely,
}
} Narahari.
}
} ________________________________________________________________________________
}
} Narahari Ramanuja
}
} Office : Materials Synthesis & Characterization Lab,
} Dept of Materials Science & Engineering,
} NJIT,
} ph # 201-596-3680
} email : nxr0308-at-megahertz.njit.edu
} ________________________________________________________________________________
}





From: Larry Allard :      l2a-at-ornl.gov
Date: Thu, 03 Dec 1998 12:44:37 -0500
Subject: Re: Centaurus BSD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Russ:

Try http://www.gwelectronics.com/

Larry






} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Dr. Lawrence F. Allard
Senior Research Staff Member
High Temperature Materials Laboratory
Oak Ridge National Laboratory
1 Bethel Valley Road
Bldg. 4515, MS 6064
PO Box 2008
Oak Ridge, TN 37831-6064

423-574-4981
423-574-4913 Fax
l2a-at-ornl.gov







From: Kim Rensing :      Kim_Rensing-at-bc.sympatico.ca
Date: Thu, 3 Dec 1998 09:54:43 -0800
Subject: glow discharge and methyl formiate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

I forwarded some recent postings on negative staining for bacteriophages to
a colleague. He is now interested in some details on the techniques for
glow discharge and on the benefits or drawbacks of using uranyl formiate
instead of uranyl acetate. If anyone could post some references or personal
experiences, both he and I would be grateful.

Thank you.

Kim





From: Bernard Kestel :      bernard_kestel-at-qmgate.anl.gov
Date: 3 Dec 1998 13:27:59 -0600
Subject: Re: NiTi etchant

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
"microscopy-at-sparc5.microscopy.co" {microscopy-at-Sparc5.Microscopy.Com}
X-Mailer: Mail*Link SMTP-QM 4.1.0



From: Barry, Lilith :      Lilith.Barry-at-nrc.ca
Date: Thu, 3 Dec 1998 14:48:00 -0500
Subject: Antibody to cy3

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America

12/3/98 10:34 AM

From a Leco Corp. metallography book-1977
3000 lakeview Ave., St. Joseph, Michigan, 49085

#50 5ml acetic acid Electolytic, platinum touch wire at 1.5 volts
10ml nitric acid for 20 to 60 seconds
85ml water

# 133 50ml nitric acid Mix fresh. Swab 5 to 30 seconds or electolytic at
50ml acetic acid 5-10 volts, for 5 to 60 seconds.
(Polishes at high currents).

#143 0.01-1 g Allow mixture to age a few minutes, then swab or
chromium trioxide immerse a few seconds to a few minutes.
100ml HCL

# 151 10ml HF Swab 5 to 30 seconds
25ml nitric acid (Caution: Wear gloves rated for HF acid
150ml water protection-and use a fume hood).

Note: I have not personally used any of these etches but hope the information
helps Edoardo Bemporad and others.

Bernard Kestel
Materials Science Division, Bldg. 212
Argonne National Laboratory
9700 South Cass Avenue
Argonne Illinois, 60439

E-mail: {kestel-at-anl.gov} FAX: (630) 252-4289


--------------------------------------

Any suggestion on proper etchant procedure to use in order to do NiTi
Memory Shape alloys SEM images?
Thank You in advance, Edoardo.



******************* NOTE *******************
There may be important message content
contained in the following MIME Information.
********************************************


------------------ MIME Information follows ------------------

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


This message is in MIME format. Since your mail reader does not understand
this format, some or all of this message may not be legible.

------ =_NextPart_001_01BE1E1D.87310D90
Content-Type: text/plain

{ { { { { { See above "Message Body" } } } } } }

------ =_NextPart_001_01BE1E1D.87310D90
Content-Type: text/html

{!DOCTYPE HTML PUBLIC "-//W3C//DTD HTML 3.2//EN"}
{HTML}
{HEAD}
{META HTTP-EQUIV="Content-Type" CONTENT="text/html; charset=us-ascii"}
{META NAME="Generator" CONTENT="MS Exchange Server version 5.5.1960.3"}
{TITLE} NiTi etchant {/TITLE}
{/HEAD}
{BODY}

{P} {FONT SIZE=2 FACE="Arial"} Any suggestion on proper etchant procedure to use in order to do NiTi Memory Shape alloys SEM images? {/FONT}
{BR} {FONT SIZE=2 FACE="Arial"} Thank You in advance, Edoardo. {/FONT}
{/P}

{/BODY}
{/HTML}
------ =_NextPart_001_01BE1E1D.87310D90--


------------------ RFC822 Header Follows ------------------
Received: by qmreceive.anl.gov with ADMIN;2 Dec 1998 12:06:39 -0600
Received: from Sparc5.Microscopy.Com (sparc5.microscopy.com [206.69.208.10]) by dns2.anl.gov (8.8.7/8.6.11) with SMTP id MAA12983; Wed, 2 Dec 1998 12:05:15 -0600 (CST)
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11) id LAA10110 for dist-Microscopy; Wed, 2 Dec 1998 11:41:38 -0600
Received: from no_more_spam.com (Sparc5 [206.69.208.10]) by Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id LAA10104 for "MicroscopyFilteredEmail-at-msa.microscopy.com"; Wed, 2 Dec 1998 11:41:07 -0600
Received: from materials10.dimi.uniroma3.it (materials10.dimi.uniroma3.it [193.205.140.46]) by Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id LAA10087 for {microscopy-at-Sparc5.Microscopy.Com} ; Wed, 2 Dec 1998 11:40:54 -0600
Received: by materials10.dimi.uniroma3.it with Internet Mail Service (5.5.1960.3)
id {YD9XW0QD} ; Wed, 2 Dec 1998 18:59:42 +0100
Message-ID: {E11E487CBB27D2118BD400A0C9DF502C37DE-at-materials10.dimi.uniroma3.it}


Would someone know if there is antibody to cy3 and who carries it. I know
that there are few companies that sell antibodies for FITC.
Internet search gives me all the cy3 conjugates.
Thank you,
Lilith
---------------------
Lilith Ohannessian-Barry
NRC, IBS,
Ottawa, Ont. K1A 0R6
CANADA
Tel;613-993-6460
Fax;613-941-4475
e-mail; lilith.barry-at-nrc.ca





From: Tom Phillips :      tphillips-at-biosci.mbp.missouri.edu
Date: Thu, 3 Dec 1998 14:25:53 -0600
Subject: Re: glow discharge and methyl formiate / neg staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


--============_-1299405738==_ma============
Content-Type: text/plain; charset="us-ascii"


}
} Hello all,
}
} I forwarded some recent postings on negative staining for bacteriophages to
} a colleague. He is now interested in some details on the techniques for
} glow discharge and on the benefits or drawbacks of using uranyl formiate
} instead of uranyl acetate. If anyone could post some references or personal
} experiences, both he and I would be grateful.
}
} Thank you.
}
} Kim

The original thread on negative staining bacteriophages was in response to
a query I made. I am appending to this message a slightly condensed
summary of all the replies. I haven't tried all the suggestions yet but I
did get much better staining from Ammonium Hydroxide neutralized PTA. The
Am Molybdate was different and needs further examination. I plan to try
additional suggestions as soon as I get a fresh phage prep. The advice
once again proved how useful this listserver is to me. Thanks to everybody
who shared their experiences. Tom

In doing Bacteria, we found it advantagous to dilute the stain 1/30 with
water. This made the Bacteria show up with good inner structure instead of
"the blob".

Perhaps a more dilute solution may stain a little less, but show up fine
detail better? If that doesn't work... more concentrated? ** Also did you
fix the pacterophages? Fixing may help
***************************
Lou Ann Miller Microscopic Imaging Lab College of Vet. Medicine U
of Illinois
2001 S Lincoln Ave Urbana,Illinois 61801 217-244-1566
lamiller-at-ux1.cso.uiuc.edu

_____________________________________________________________________

a) Did you 'glow discharge' the grid shortly before you adsorbed the drop?
b) Have you tried Uranyl Formiate solution? (I think 0.75%w/w and pH around
4 should be good.)
c) Have you tried lower concentration of Uranyl salt? Maybe the fine
details are embedded in a thick salt layer so that you don't get any
contrast there.

The procedure I normally use is:
- prepare parafilm with at least 2 droplets of water (or sometimes buffer?) and
another two of stain solution
- glow discharge the grid for around 10 to 20 s (blue-white)
- clamp it into tweezer
- put 5um drop on the grid and wait... whatever.. 30 s, 60 s....
- touch with filter paper to soak solution away (blotting)
- wash on first drop of water by touching the surface, blot again
- repeat above step at least once
- then go to stain droplet and do the same
- the second stain droplet is then used to do the staining procedure, so touch
the surface and wait for a few seconds (lets say 10 to 20)
- blot a last time and that's it

You may not need all these steps, but I have been very successful with this
technique and don't see any reason not to do it this way. If you don't have
Uranyl Formiate, you can try the upper protocol with Uranyl Acetate (2% or
lower). My experience is that usually it should work as nicely as UFo, but
I was told that UAc does 'melt' more strongly and more quickly in the
electron beam. I am looking forward to other recipies. I am very sure that
you can find as many recipies as there are microscopists...

By the way, I have never been successful using PTA. Even with very
sensitive proteins which don't like a low pH, I always had the best results
in using Uranyl solutions.

Bettina Wolpensinger Electron Microscope Unit University of New
South Wales
Sydney, NSW 2052, AUSTRALIA b.wolpensinger-at-unsw.edu.au
http://srv.emunit.unsw.edu.au

_____________________________________________________________________

I'm no expert but we have had problems because of the following:

My experience with UA is that when it works it's great but it doesn't
always work.

With PTA a bad batch of PTA stock could result in poor images - especially
if it's old or not really a good quality reagent..

Do you adjust the pH of your PTA? It is normally fairly acid and should be
adjusted to between 6 and 7 but you can experiment. We have always
understood that it is best to use potassium hydroxide and not sodium
hydroxide to adjust - personally I've seen little difference.

If you are using carbon coatings on your grids and clean preparations of
virus maybe you need a little BSA to help spreading (not normally a problem
with our samples).

If none of the above work it might be worth trying some of the newer more
expensive stains like sodium silicotungstate or methylamine tungstate which
can give nice results.

I suspect you may have tried most of the above, but I hope that something
will help. Good Luck.

Malcolm Haswell Electron Microscopy School of Health Sciences
Fleming Bldg
University of Sunderland SUNDERLAND SR1 3SD Tyne and Wear UK
Tel (0191) 515 2872 e-mail: malcolm.haswell-at-sunderland.ac.uk
_____________________________________________________________________

Several years a go our lab switched from the traditional KPTA to NHPTA.
which enhances the fine structural details often lost with more aggressive
stains. Prepare a 1% solution of phosphotungstic acid , then pH the stain
with 1N ammonium hydroxide to 6.5 and / or 7.0.
Regards, Skip
} From: L R MELSEN {lmelsen-at-emory.edu}
_____________________________________________________________________
When you stain (this is for t4 and UAC) allow four to eight drops of stain
I use, almost as routine, ammonium molibdate, 1 or 2% in water with very
good results (for visualize African swine fever virus, herpesvirus,
rotaviruses, polyomaviruses, adenoviruses). My protocol: (1) a drop of the
virus containing pellet over a carbon coated grid; (2) wash in situ with
5-7 drops of negative stain; (3) dry with paper filter.

J.F.Moura Nunes Lab. Electron Microscopy
Portuguese Cancer Institute 1093 Lisbon Codex Portugal

_____________________________________________________________________
When you stain (this is for t4 and UAC) allow four to eight drops of stain
to drip on and off the grid. This works well if you hold the grid at a 45
degree angle. If you wash try 0.1mNaCl. Water can osmotically shock the
specimen.
Bob
BOB-at-befvax.uchicago.edu


____________________________________________________________________
I would only add to Paul Webster's and other's comments that an amorphous
carbon film that has been glow-discharged in a partial vacuum with air for
about 60 seconds is pretty hard to beat when it comes to a substrate for
negative staining. We used this procedure for T4 on one of my (numerous)
postdocs in the Dept. of Microbiology at the Biozentrum in Basel. We never
found anything to beat it. The stain spreads beautifully on the surface.

Robert (Bob) Chiovetti Microimaging Technologies, Inc.
Tucson, Arizona USA Tel. / Fax (520) 546-4986
rchiovetti-at-aol.com

_____________________________________________________________________
In addition to the many recipes for stains you must be receiving, you
might also like to try a simple "low-dose" imaging protocol. I have found
when using negative stains that the image selection and focussing
procedures can damage fragile structures in the sample. Focussing on an
area close to a particle and then moving over to take the picture can help
prevent this. If you own a newer TEM, you might even have a low dose
capability built in. I have found this method very useful for small,
fragile samples, even when coated with thick films of aqueous uranyl
acetate or phosphotungstic acid.

Alternatively, the support film might be too thick to image the fine
detail. Is your grid carbon coated, as you post, or is it formvar/carbon
coated, which is more usual? Carbon alone offers more for high resolution
work. Finally, have you checked out the vibrations affecting the
microscope? High resolution requires stable conditions. Check out my web
site for a brief summary of the options {http://www.hei.org/htm/neg.htm} .
I look forward to postings containing stain receipies.

Paul Webster, Ph.D House Ear Institute 2100 West Third Street Los
Angeles, CA 90057 phone:213 273 8026 fax: 213 413 6739
e-mail: pwebster-at-hei.org
http://www.hei.org/htm/aemi.htm

_____________________________________________________________________
Try 3% ammonium molybdate pH 7.2 , I have good results with it.

Ann Fook Yang EM Unit Eastern Cereal and Oilseed Research Centre
Agriculture and Agri-Food Canada 960 Carling Ave Central
Experimental Farm
Ottawa, Ontario Canada K1A 0C6 Tel.: 613-759-1638 Fax: 613-759-1701

e-mail:yanga-at-em.agr.ca

_____________________________________________________________________
Look up work by this expert:

Ackermann HW. Krisch HM.
Felix d'Herelle Reference Center for Bacterial Viruses, Department
of Microbiology, Faculty of Medicine, Laval University, Sainte-Foy, Canada.
A catalogue of T4-type bacteriophages. [Review] [115 refs]
Archives of Virology. 142(12):2329-45, 1997.

Sara E. Miller, Ph. D. P. O. Box 3020 Duke University Medical
Center Durham, NC 27710 Ph: 919 684-3452 FAX: 919 684-8735
Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility

3 Tucker Hall
Division of Biological Sciences
University of Missouri
Columbia, MO 65211
(573)-882-4712 (voice)
(573)-882-0123 (fax)
--============_-1299405738==_ma============
Content-Type: text/enriched; charset="us-ascii"



}

} Hello all,

}

} I forwarded some recent postings on negative staining for
bacteriophages to

} a colleague. He is now interested in some details on the techniques
for

} glow discharge and on the benefits or drawbacks of using uranyl
formiate

} instead of uranyl acetate. If anyone could post some references or
personal

} experiences, both he and I would be grateful.

}

} Thank you.

}

} Kim


The original thread on negative staining bacteriophages was in response
to a query I made. I am appending to this message a slightly condensed
summary of all the replies. I haven't tried all the suggestions yet
but I did get much better staining from Ammonium Hydroxide neutralized
PTA. The Am Molybdate was different and needs further examination. I
plan to try additional suggestions as soon as I get a fresh phage prep.
The advice once again proved how useful this listserver is to me.
Thanks to everybody who shared their experiences. Tom


In doing Bacteria, we found it advantagous to dilute the stain 1/30
with water. This made the Bacteria show up with good inner structure
instead of "the blob".


Perhaps a more dilute solution may stain a little less, but show up
fine detail better? If that doesn't work... more concentrated? **
Also did you fix the pacterophages? Fixing may help

***************************

Lou Ann Miller Microscopic Imaging Lab College of Vet. Medicine U of
Illinois

2001 S Lincoln Ave Urbana,Illinois
61801 217-244-1566 lamiller-at-ux1.cso.uiuc.edu


_____________________________________________________________________


a) Did you 'glow discharge' the grid shortly before you adsorbed the
drop?

b) Have you tried Uranyl Formiate solution? (I think 0.75%w/w and pH
around 4 should be good.)

c) Have you tried lower concentration of Uranyl salt? Maybe the fine
details are embedded in a thick salt layer so that you don't get any
contrast there.


The procedure I normally use is:

- prepare parafilm with at least 2 droplets of water (or sometimes
buffer?) and

another two of stain solution

- glow discharge the grid for around 10 to 20 s (blue-white)

- clamp it into tweezer

- put 5um drop on the grid and wait... whatever.. 30 s, 60 s....

- touch with filter paper to soak solution away (blotting)

- wash on first drop of water by touching the surface, blot again

- repeat above step at least once

- then go to stain droplet and do the same

- the second stain droplet is then used to do the staining procedure,
so touch

the surface and wait for a few seconds (lets say 10 to 20)

- blot a last time and that's it


You may not need all these steps, but I have been very successful with
this technique and don't see any reason not to do it this way. If you
don't have Uranyl Formiate, you can try the upper protocol with Uranyl
Acetate (2% or lower). My experience is that usually it should work as
nicely as UFo, but I was told that UAc does 'melt' more strongly and
more quickly in the electron beam. I am looking forward to other
recipies. I am very sure that you can find as many recipies as there
are microscopists...


By the way, I have never been successful using PTA. Even with very
sensitive proteins which don't like a low pH, I always had the best
results in using Uranyl solutions.


Bettina Wolpensinger Electron Microscope Unit University of New South
Wales

Sydney, NSW 2052, AUSTRALIA b.wolpensinger-at-unsw.edu.au
http://srv.emunit.unsw.edu.au


_____________________________________________________________________


I'm no expert but we have had problems because of the following:


My experience with UA is that when it works it's great but it doesn't
always work.


With PTA a bad batch of PTA stock could result in poor images -
especially if it's old or not really a good quality reagent..


Do you adjust the pH of your PTA? It is normally fairly acid and should
be

adjusted to between 6 and 7 but you can experiment. We have always
understood that it is best to use potassium hydroxide and not sodium
hydroxide to adjust - personally I've seen little difference.


If you are using carbon coatings on your grids and clean preparations
of virus maybe you need a little BSA to help spreading (not normally a
problem with our samples).


If none of the above work it might be worth trying some of the newer
more expensive stains like sodium silicotungstate or methylamine
tungstate which can give nice results.


I suspect you may have tried most of the above, but I hope that
something will help. Good Luck.


Malcolm Haswell Electron Microscopy School of Health Sciences Fleming
Bldg

University of Sunderland SUNDERLAND SR1 3SD Tyne and Wear UK

Tel (0191) 515 2872 e-mail:
malcolm.haswell-at-sunderland.ac.uk

_____________________________________________________________________


Several years a go our lab switched from the traditional KPTA to NHPTA.
which enhances the fine structural details often lost with more
aggressive stains. Prepare a 1% solution of phosphotungstic acid , then
pH the stain with 1N ammonium hydroxide to 6.5 and / or 7.0.

Regards, Skip

} From: L R MELSEN { {lmelsen-at-emory.edu}

_____________________________________________________________________

When you stain (this is for t4 and UAC) allow four to eight drops of
stain I use, almost as routine, ammonium molibdate, 1 or 2% in water
with very good results (for visualize African swine fever virus,
herpesvirus, rotaviruses, polyomaviruses, adenoviruses). My protocol:
(1) a drop of the virus containing pellet over a carbon coated grid;
(2) wash in situ with 5-7 drops of negative stain; (3) dry with paper
filter.


J.F.Moura Nunes Lab. Electron Microscopy

Portuguese Cancer Institute 1093 Lisbon Codex Portugal


_____________________________________________________________________

When you stain (this is for t4 and UAC) allow four to eight drops of
stain to drip on and off the grid. This works well if you hold the
grid at a 45 degree angle. If you wash try 0.1mNaCl. Water can
osmotically shock the specimen.

Bob

BOB-at-befvax.uchicago.edu



____________________________________________________________________

I would only add to Paul Webster's and other's comments that an
amorphous

carbon film that has been glow-discharged in a partial vacuum with air
for

about 60 seconds is pretty hard to beat when it comes to a substrate
for

negative staining. We used this procedure for T4 on one of my
(numerous)

postdocs in the Dept. of Microbiology at the Biozentrum in Basel. We
never

found anything to beat it. The stain spreads beautifully on the
surface.


Robert (Bob) Chiovetti Microimaging Technologies, Inc.

Tucson, Arizona USA Tel. / Fax (520) 546-4986

rchiovetti-at-aol.com


_____________________________________________________________________

In addition to the many recipes for stains you must be receiving, you
might also like to try a simple "low-dose" imaging protocol. I have
found when using negative stains that the image selection and focussing
procedures can damage fragile structures in the sample. Focussing on
an area close to a particle and then moving over to take the picture
can help prevent this. If you own a newer TEM, you might even have a
low dose capability built in. I have found this method very useful for
small, fragile samples, even when coated with thick films of aqueous
uranyl acetate or phosphotungstic acid.


Alternatively, the support film might be too thick to image the fine
detail. Is your grid carbon coated, as you post, or is it
formvar/carbon coated, which is more usual? Carbon alone offers more
for high resolution work. Finally, have you checked out the vibrations
affecting the microscope? High resolution requires stable conditions.
Check out my web site for a brief summary of the options
{ {http://www.hei.org/htm/neg.htm} . I look forward to postings
containing stain receipies.


Paul Webster, Ph.D House Ear Institute 2100 West Third Street Los
Angeles, CA 90057 phone:213 273 8026 fax: 213 413 6739 e-mail:
pwebster-at-hei.org

http://www.hei.org/htm/aemi.htm


_____________________________________________________________________

Try 3% ammonium molybdate pH 7.2 , I have good results with it.


Ann Fook Yang EM Unit Eastern Cereal and Oilseed Research Centre

Agriculture and Agri-Food Canada 960 Carling Ave Central Experimental
Farm

Ottawa, Ontario Canada K1A 0C6 Tel.: 613-759-1638 Fax: 613-759-1701

e-mail:yanga-at-em.agr.ca


_____________________________________________________________________

Look up work by this expert:


Ackermann HW. Krisch HM.

Felix d'Herelle Reference Center for Bacterial Viruses, Department

of Microbiology, Faculty of Medicine, Laval University, Sainte-Foy,
Canada.

A catalogue of T4-type bacteriophages. [Review] [115 refs]


Archives of Virology. 142(12):2329-45, 1997.


Sara E. Miller, Ph. D. P. O. Box 3020 Duke University Medical
Center Durham, NC 27710 Ph: 919 684-3452 FAX: 919 684-8735

Thomas E. Phillips, Ph.D.

Associate Professor of Biological Sciences

Director, Molecular Cytology Core Facility


3 Tucker Hall

Division of Biological Sciences

University of Missouri

Columbia, MO 65211

(573)-882-4712 (voice)

(573)-882-0123 (fax)

--============_-1299405738==_ma============--





From: David Smith :      idleweed-at-tusco.net
Date: Thu, 3 Dec 1998 16:05:27 -0500
Subject: need info for scanning insects

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.

------=_NextPart_000_0029_01BE1ED6.BDE3A540
Content-Type: text/plain;
charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

I understand that there is information on scanning and digital cameras =
at this address. Is that true and how do I access it

------=_NextPart_000_0029_01BE1ED6.BDE3A540
Content-Type: text/html;
charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

{!DOCTYPE HTML PUBLIC "-//W3C//DTD W3 HTML//EN"}
{HTML}
{HEAD}

{META content=3Dtext/html;charset=3Diso-8859-1 =
http-equiv=3DContent-Type}
{META content=3D'"MSHTML 4.72.3110.7"' name=3DGENERATOR}
{/HEAD}
{BODY bgColor=3D#ffffff}
{DIV} {FONT color=3D#000000 size=3D2} I understand that there is =
information on=20
scanning and digital cameras at this address. Is that true and how do I =
access=20
it {/FONT} {/DIV} {/BODY} {/HTML}

------=_NextPart_000_0029_01BE1ED6.BDE3A540--






From: Russell E. Cook :      cook-at-aaem.amc.anl.gov
Date: Thu, 3 Dec 1998 15:08:11 -0600
Subject: re Centaurus

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to all who replied. For those of you who might want the
information, the Centaurus back-scattered electron detector can be found in
the USA at:

http://www.gwelectronics.com/backed.htm

GW Electronics Inc.
6981 Peachtree Industrial Blvd.
Norcross, GA 30092-3601
Tel: 770-449-0707
800-325-5556

Mark Massey of EDAX Europe told me that they are manufactured by KE
Developments in the UK:
K. E. Development Limited
The Mount, Toft
Cambridge CB3 7RL
UK
Tel: +44 1223 263 948
Fax: +44 1223 263 532

Russell E. Cook
Electron Microscopy Center for Materials Research
Argonne National Laboratory
Building 212
9700 South Cass Avenue
Argonne, IL 60439
(630)252-7194
FAX: (630)252-4798







From: GANTZ-at-med-biophd.bu.edu
Date: Thu, 03 Dec 1998 17:57:51 -0400 (EDT)
Subject: Glow Discharge

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


NiTi alloys are very sensitive to structural damage during polishing. Be
sure to use care in mechanical polishing or use electro-polishing.

Electro-polish: 20% H2SO4 in MeOH 28-29v for 1-5 min.

Immersion etch: similar to ASTM etch No. 151 and ASTM etch No. 209
14.1ml HNO3
3.2ml HF
82.7ml H2O
submerge for 10-15 sec. with gentile agitation.

Good luck
Sam. O. Mancuso
Group Leader,
Physical-Mechanical Metallurgy
Special Metals Corporation
4317 Middle Settlement Road
New Hartford, New York 13413
U.S.A.
{} {} {}
Phone: (315) 798-2920
Fax: (315) 798-2001
email: mancuso4-at-ix.netcom.com

-----Original Message-----
} From: Edoardo Bemporad [mailto:e.bemporad-at-materials10.dimi.uniroma3.it]
Sent: Wednesday, December 02, 1998 9:56 AM
To: microscopy-at-sparc5.microscopy.com


Dear Kim:
An excellent reference on glow discharging is in Advances in Optical
and Electron Microscopy 8:107-135 (1982) by Dubochet, Groom and Mueller-
Neuteboom.
Don Gantz
Boston Univ Med School
gantz-at-med-biophd.bu.edu





From: Stephen R Poe :      Stephen.R.Poe-at-usda.gov
Date: Thu, 3 Dec 1998 17:40:09 -0500
Subject: TEM - LKB Ultrotome III sale/trade, Help with LKB knife breaker

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


OK, well I have really done it now - bought a whole TEM sectioning workstation
in order to get the glass knife breaker. I am an LM type who is starting to
use glass knives for some of my sectioning work (that is another story).
Anyway, this has created two problems:

1. I need a copy of the manual for an LKB 7801-B knife breaker so I can get
going with this project. I think I remember how from my grad school days, but
would be a lot more confident with a copy of the manual. Would be happy to pay
for the copying and postage charges.

2. I am left with an LKB Ultrotome III in what is reported to be good working
condition - this is surplus to my rather simple needs (I am currently doing my
sectioning on a nice old Lipshaw microtome). The microtome consists of the
mail unit (8800), the control unit (8802-A), the base (8809-A), and the
illuminator base for the sterio microscope head (no sterio microscope of
course). Again, I have not used this unit, so I can not vouch for it's
mechanical/operating condition - it is clean and apparently complete. I would
be open to reasonable offers - biased in favor of someone in the Washington DC
- Baltimore, MD area. This thing is HEAVY, and I would have to be pretty well
compensated to want to build a proper crate for it. It would be much easier
for someone to pick it up, or at least if I could arrange to meet someone half
way. As they say, money is always nice, but interesting trades for LM
equipment are also quite possible (my lab is based on old Leitz 170mm Ortholux
and Laborlux scopes, and a variety of AO and B&L sterio scopes - those I need
more of always).

Thanks for any assistance.

Stephen Poe - I can also be reached at my at-home E-mail address
spoefish-at-mindspring.com





From: Angela Klaus :      avklaus-at-amnh.org
Date: Thu, 03 Dec 1998 18:11:10 -0500
Subject: Re: Centaurus BSD

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Russell:

The Centaurus BSE detector is marketed by GW Electronics, (770)449-0707.

Best regards,

Angela

} Who manufacturesand/or markets the Centaurus BSE detector?
}
} Russell E. Cook
} Electron Microscopy Center for Materials Research
} Argonne National Laboratory
} Building 212
} 9700 South Cass Avenue
} Argonne, IL 60439
} (630)252-7194
} FAX: (630)252-4798
}
---------------------------------------------
Angela V. Klaus

Manager - Core Microscopy Facility
American Museum of Natural History
Central Park West at 79th Street
New York, NY 10024-5192 USA

Email: avklaus-at-amnh.org
Voice: (212)769-5977, 5469
Fax: (212)769-5495
---------------------------------------------





From: Stephen R Poe :      Stephen.R.Poe-at-usda.gov
Date: Thu, 3 Dec 1998 17:44:30 -0500
Subject: TEM - LKB Ultrotome III sale/trade, Help with LKB knife breaker

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


OK, well I have really done it now - bought a whole TEM
sectioning workstation in order to get the glass knife breaker. I am an LM
type who is starting to use glass knives for some of my sectioning work (that
is another story). Anyway, this has created two problems:

1. I need a copy of the manual for an LKB 7801-B knife breaker so I can get
going with this project. I think I remember how from my grad school days, but
would be a lot more confident with a copy of the manual. Would be happy to pay
for the copying and postage charges.

2. I am left with an LKB Ultrotome III in what is reported to be good working
condition - this is surplus to my rather simple needs (I am currently doing my
sectioning on a nice old Lipshaw microtome). The microtome consists of the
mail unit (8800), the control unit (8802-A), the base (8809-A), and the
illuminator base for the sterio microscope head (no sterio microscope of
course). Again, I have not used this unit, so I can not vouch for it's
mechanical/operating condition - it is clean and apparently complete. I would
be open to reasonable offers - biased in favor of someone in the Washington DC
- Baltimore, MD area. This thing is HEAVY, and I would have to be pretty well
compensated to want to build a proper crate for it. It would be much easier
for someone to pick it up, or at least if I could arrange to meet someone half
way. As they say, money is always nice, but interesting trades for LM
equipment are also quite possible (my lab is based on old Leitz 170mm Ortholux
and Laborlux scopes, and a variety of AO and B&L sterio scopes - those I need
more of always).

Thanks for any assistance.

Stephen Poe - I can also be reached at my at-home E-mail address
spoefish-at-mindspring.com





From: Stephen R Poe :      Stephen.R.Poe-at-usda.gov
Date: Thu, 3 Dec 1998 17:53:16 -0500
Subject: TEM - LKB Ultrotome III sale/trade, help with LKB knife breaker

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


OK, well I have really done it now - bought a whole TEM sectioning workstation
in order to get the glass knife breaker. I am an LM type who is starting to
use glass knives for some of my sectioning work (that is another story).
Anyway, this has created two problems:

1. I need a copy of the manual for an LKB 7801-B knife breaker so I can get
going with this project. I think I remember how from my grad school days, but
would be a lot more confident with a copy of the manual. Would be happy to pay
for the copying and postage charges.

2. I am left with an LKB Ultrotome III in what is reported to be good working
condition - this is surplus to my rather simple needs (I am currently doing my
sectioning on a nice old Lipshaw microtome). The microtome consists of the
mail unit (8800), the control unit (8802-A), the base (8809-A), and the
illuminator base for the sterio microscope head (no sterio microscope of
course). Again, I have not used this unit, so I can not vouch for it's
mechanical/operating condition - it is clean and apparently complete. I would
be open to reasonable offers - biased in favor of someone in the Washington DC
- Baltimore, MD area. This thing is HEAVY, and I would have to be pretty well
compensated to want to build a proper crate for it. It would be much easier
for someone to pick it up, or at least if I could arrange to meet someone half
way. As they say, money is always nice, but interesting trades for LM
equipment are also quite possible (my lab is based on old Leitz 170mm Ortholux
and Laborlux scopes, and a variety of AO and B&L sterio scopes - those I need
more of always).

Thanks for any assistance.

Stephen Poe - I can also be reached at my at-home E-mail address
spoefish-at-mindspring.com





From: Ronald Capel :      r.capel-at-student.utwente.nl
Date: Fri, 04 Dec 1998 01:26:36 +0100
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html









From: McLean, Dorrance :      dmclea-at-sandia.gov
Date: Thu, 3 Dec 1998 18:23:46 -0700
Subject: Al-Si alloy-Electrolytic thinning-HELP!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

I've been batting my head against my Fischione Jet Polisher trying to figure
this one out! Please help if you can. My original material is full of
lovely voids and pits, Si particles are around 10 microns, and at a
thickness of 150 to 200 microns I have experienced some holes that go all
the way through the material. If I try initial thinning of the sample on
less than 500 grit SiC paper I get even more particle pullout...Any ideas?

Thanks in advance for any help.

Dorrance





From: Meredith Wallwork :      mwallwor-at-waite.adelaide.edu.au
Date: Fri, 4 Dec 1998 12:00:35 +1030 (CST)
Subject: re: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html








From: Mike Wombwell :      mwombwell-at-vgscientific.com
Date: Thu, 3 Dec 1998 19:45:45 -0500
Subject: sputter coater

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Marty Reed Wrote:

} Our sputter coater (ISI PS-2) just died. I need parts for it and was
} hoping that someone may have an old one setting around that I could
} get for parts. I would also consider a used sputter coater of any
} make that you may want to sell.

} Thank you

} Marty Reed
} Equipment Technician
} Biology Department
} Humboldt State University


Dear Marty,

The ISI PS-2 sputter coater was made by us way back in the early
1970s during our days as Polaron Equipment Ltd. Although the
PS-2 has not been manufactured for many years we do still carry some
spare parts.
For further information please contact our local distributor, Energy
Beam Sciences Inc. EBS ebs-at-ebsciences.com tel: 413 786 9322

Please note our new address, telephone and fax number for any
enquiries concerning the Polaron range


Best regards
Mike Wombwell
Polaron range Business Manager
V G Microtech
The Birches Industrial Estate
Imberhorne Lane
East Grinstead
West Sussex
RH19 1UB
UK
Direct line: +44 (0)1342 310296
Switchboard: +44 (0)1342 327211
Fax: +44 (0)1342 315074
http://www.vgmicrotech.com/polaron-range
E&OE







From: Amanda Harman :      biajh-at-CC.newcastle.edu.au
Date: Thu, 3 Dec 1998 22:26:38 -0500
Subject: Permanent fluorescent mounting media

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would be grateful if anyone could tell me if there is a fluorescent
mounting media that is permanent?
Amanda Harman.







From: Stephan Coetzee :      stephan-at-gecko.biol.wits.ac.za
Date: Fri, 4 Dec 1998 08:36:41 GMT+2
Subject: Re: Al-Si alloy-Electrolytic thinning-HELP!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dorrance
Before I will be able to suggest some advice, I would need more
information.


} My original material is full of
} lovely voids and pits, Si particles are around 10 microns, and at a
} thickness of 150 to 200 microns I have experienced some holes that go all
} the way through the material.

Firstly what is your original material? That will determine quite a
few tings Eg. rate of thinning vs Si particle thinning rate.

} If I try initial thinning of the sample on
} less than 500 grit SiC paper I get even more particle pullout...Any ideas?
How is your sample "hold' in order to polish. Is it already 3mm
diameter or will you punch the disks later. what lubrication do you
use if any.
}
} Thanks in advance for any help.
Mr. S H Coetzee Tell: (011) 716 2419
Electron Microscope Unit Fax: (011) 339 3407
Private bag X3 E-mail: Stephan-at-gecko.biol.wits.ac.za
Wits
Johannesburg
2050





From: Jacob Bastacky :      sjbastacky-at-lbl.gov
Date: Thu, 3 Dec 1998 23:03:06 -0800
Subject: Query: How much better: WDS than EDS for SODIUM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are considering looking for a Wavelength-Dispersive Spectrometer to
improve our detection of sodium in biological samples, reasoning that the
better energy resolution of the detector will result in narrower, taller
peaks with the same background level, yielding improved peak-to-background
ratios.

Our samples are frozen hydrated and dried biological tissues; we would like
to be able to differentiate 50mMolar from 150mMolar sodium (0.25 and .75%
by weight in hydrated samples).

Our present EDS detector has a semi-thin window and is doing better than
average for sodium and we would prefer not going to an ultrathin window or
windowless detector.

Might we expect a significant improvement with WDS?

Does anyone know of a spectrometer that might be donated?

Your comments and suggestions will be appreciated.

Thanks,

Jacob

Jacob Bastacky, M.D.
Room 116 Donner Laboratory
Lawrence Berkeley Laboratory
University of California
Berkeley, California 94720
Telephones: 510.486.4606 office, 510.486.4605 lab, 510.845.8031 fax
email: sjbastacky-at-lbl.gov







From: Guenter Giese :      giese-at-pluto.mpi-hd.mpg.de
Date: Fri, 04 Dec 1998 09:07:37 +0100
Subject: Printouts of RGB fluorescence images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I am sure that I am not the only one who runs into problems with color
representation when printing out fluorescence (RGB) images on paper. I
think that I am aware of the cause for the problem, but I got no final
solution for it.

While fluorescence images (and other images representing brightness values
as pure red, green or blue colors) look pretty well on the monitor or when
presented as a slide or with a laser beamer (working in RGB mode), the
translation of channels into colors renders pure colors and most color
combinations too dark on the printout. This color translation has, in
principle, nothing to do with the translation of RGB colors into YMCK
printer colors by the printer driver, but is rather a consequence of a
one-to-one translation of channels into pure colors. For instance, printing
pixels representing bright green fluorescence (e.g. the maximum brightness
value of 255 at 8 bit resolution per channel) results in a saturated green
(maximum green value) on the printout, which appears too dark to the eye.
Remembering bright green fluorescence in the microscope, our eyes await to
see a bright green color (white/green) on the printout, which may be
represented by a certain mixture of green, red and blue (or the respective
mixture of CMYK printer colors).

I think what would help would be a good look-up table or a set of tables
(or, e.g., a Photoshop macro) to represent fluorescence images (single
channels and channel combinations) appropriately on different color printer
types (laser, inkjet and thermotransfer/thermoosublimation printers).
Experimenting with Photoshop (and other software's) parameters (brightness
and saturation corrections etc.) did not result in a color representation I
think (and dream) of, and is error-prone.

I am aware of the fact that, in general, printers cannot represent the
whole RGB color space (especially green colors), but I think that good
look-up tables or macros could help a lot.

Thanks in advance

Guenter



----------------------------------
Dr. Guenter Giese
MPI fuer Medizinische Forschung
Jahnstr. 29
D-69120 Heidelberg
Phone (Germany or 0-)6221-486-320
Fax (Germany or 0-)6221-486-325
e-mail: ggiese-at-mzf.mpimf-heidelberg.mpg.de
------------------------------------------





From: Rieck, Vern :      rieck-at-aavid.com
Date: Fri, 4 Dec 1998 08:18:13 -0500
Subject: RE: fine metal tip used in microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tong, If you can use tungsten for your application, McCrone Accessories
and Components sells pre-made probes and will also sell you tungsten
wire and sodium nitrite so you can etch your own wire down to the
desired size. No financial interest in McCrone, just a satisfied
customer.

http://www.mccrone.com/cgi-bin/SoftCart.exe/mac/home.html?E+mccrone

Vern Rieck
Quality Engineer
Aavid Thermal Products, Inc.
603-629-2224 direct
603-666-4100 main
fax 603-669-2044
rieck-at-aavid.com


-----Original Message-----
From: Tong Wang [SMTP:tong-at-jlab.org]
Sent: Thursday, December 03, 1998 11:50 AM
To: microscopy-at-sparc5.microscopy.com
Subject: fine metal tip used in microscopy


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

-----------------------------------------------------------------------.


Hi,
Anybody knows where to buy some metal tips of 10 micron and 1
micron
diameter? Thanks in advance.

Tong







From: Rik Brydson :      MTLRMDB-at-LUCS-MAC.NOVELL.LEEDS.AC.UK
Date: Fri, 4 Dec 1998 14:08:50 GMT0BST
Subject: LEMAS - Electron Microscopy and Analysis Listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Dear Member,

Some information on a new listserver recently set up. I encourage you to join
Please display in your institution.

Many thanks,

Rik Brydson

**********************************************************
LEMAS
Listserver for Electron Microscopy and Analytical Spectroscopy

o Designed to encourage international discussion of all aspects of electron microscopy and
microanalytical spectroscopies in both the physical and biological sciences.

o Designed to promote international information exchange (e.g. tips, ideas, meetings,
conferences) and determination of best practice in EM and analysis.

o Designed to provide a forum for new researchers to gain expertise.

Please subscribe now and encourage your students/supervisor/colleagues to also join up.
Strength lies in the subscribers !

Nb. List is unmoderated, however, abusers and junk-mailers will be unsubscribed immediately.

Subscription Information:

Send this message to mailbase-at-mailbase.ac.uk
join lemas "firstname lastname"
but type your own personal names instead of 'firstname' and 'lastname'
E.g. join lemas Ernst Ruska
You'll then get an automatic message from the Mailbase computer,
containing a unique code. You'll have to confirm your membership by
sending back a message like this:
accept xxxx
where xxx is the code sent to you. This allows the Mailbase computer
to check your email address.

LIST-OWNER: Dr Rik Brydson,
Department Of Materials,
School of Process, Environmental and Materials Engineering,
University of Leeds,
Leeds LS2 9JT, U.K.

Tel: 0044 + (0)113 233 2369
Fax: 0044 + (0)113 242 2531
Web: http://www.leeds.ac.uk/materials

**********************************************************
_____________________________
Dr. Rik Brydson,
University Research Fellow,
Electron Optical Unit,
Department of Materials,
School of Process, Environmental and Materials Engineering
University of Leeds,
Leeds LS2 9JT,
U.K.

Tel: 44 + (0)113 233 2369
Fax: 44 + (0)113 242 2531
Web: http://www.leeds.ac.uk/materials

______________________________

**************************
Join email discussion list for all aspects of
Electron Microscopy and Analytical Spectroscopy (LEMAS)

send message "join lemas firstname lastname"
to mailbase-at-mailbase.ac.uk
**************************





From: William Tivol :      tivol-at-wadsworth.org
Date: Fri, 04 Dec 1998 08:57:57 -0500
Subject: Re: Query: How much better: WDS than EDS for SODIUM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jacob Bastacky wrote:
}
} We are considering looking for a Wavelength-Dispersive Spectrometer to
} improve our detection of sodium in biological samples, reasoning that the
} better energy resolution of the detector will result in narrower, taller
} peaks with the same background level, yielding improved peak-to-background
} ratios.

This is correct.
}
} Our samples are frozen hydrated and dried biological tissues; we would like
} to be able to differentiate 50mMolar from 150mMolar sodium (0.25 and .75%
} by weight in hydrated samples).
}
This should be pretty straight-forward.
}
} Might we expect a significant improvement with WDS?
}
Yes.

} Does anyone know of a spectrometer that might be donated?

Good luck.
Yours,
Bill Tivol





From: Robert Underwood :      underwoo-at-u.washington.edu
Date: Fri, 4 Dec 1998 08:55:04 -0800 (PST)
Subject: Re: Permanent fluorescent mounting media

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi Amanda,

One permanant media that we have used is Molecular Probes "Pro-Long
Anti-Fade. It dries hard and we store them at room temp with our
histochemically stained specimens. I don't know if I would call it
totally permanant. Samples still do fade with time and exposure.

Bob

On Thu, 3 Dec 1998, Amanda Harman wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I would be grateful if anyone could tell me if there is a fluorescent
} mounting media that is permanent?
} Amanda Harman.
}
}
}
}






From: kennedy-at-nsi.edu (Grace Kennedy)
Date: Fri, 4 Dec 1998 09:46:26 -0800
Subject: alloys

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a basic question for the materials scientists. I am interested in
finding a simple reference which might delineate the different types of
stainless steel, including relative hardness, tensile strength, etc. I'm
also interested in seeing the same information plus the electrical
characteristics of the various alloys of indium. In this case, melting
points are of interest, and the exact formulae of "Wood's metal" and
another that we use called "Cerralloy". Many thanks- Grace







From: John F. Mansfield :      jfmjfm-at-engin.umich.edu
Date: Fri, 4 Dec 1998 14:15:04 -0500
Subject: IUMAS 2000 in Kona, Hawaii!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi there, this is to notify/remind everyone that:

IUMAS 2000, The 2000 Meeting of the International Union of Microbeam=20
Analysis Societies, will be held at The King Kamehameha Hotel in=20
Kailua-Kona, HawaiiJuly 9th-13th 2000=00.

This meeting is jointly sponsored by MAS, the European MAS, The Japanese=20
MAS, the Chinese MAS, the Australian MAS and the Korean MAS. Check out the=
=20
information at:

http://www.microanalysis.org/iumas2000/iumas2000.html


Jfm.




John Mansfield PhD CPhys MInstP
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (734) 936-3352 FAX (734) 936-3352
Cellular Phone: (734) 358-7555
(Leaving a phone message at 936-3352 is preferable to 358-7555)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html
Location: Lat. 42=B0 16' 48" Long. 83=B0 43' 48"





From: Gary Liechty :      garyliechty-at-worldnet.att.net
Date: Fri, 04 Dec 1998 11:53:44 -0800
Subject: Re: Al-Si alloy-Electrolytic thinning-HELP!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------350E4D479573B0548093D7E5
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

Hi Dorrance,

Silicon Carbide abrasive is one of your main problems. You need to use diamond
for thinning. Silicon carbide does not cut silicon very well, in fact it will
fracture it and that is why you are getting pullout. Try diamond lapping film
with some glycol as the lubricant.

Good Luck,

Gary Liechty
Allied High Tech Products, Inc.

McLean, Dorrance wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Dear All,
}
} I've been batting my head against my Fischione Jet Polisher trying to figure
} this one out! Please help if you can. My original material is full of
} lovely voids and pits, Si particles are around 10 microns, and at a
} thickness of 150 to 200 microns I have experienced some holes that go all
} the way through the material. If I try initial thinning of the sample on
} less than 500 grit SiC paper I get even more particle pullout...Any ideas?
}
} Thanks in advance for any help.
}
} Dorrance



--------------350E4D479573B0548093D7E5
Content-Type: text/x-vcard; charset=us-ascii; name="vcard.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for Gary Liechty
Content-Disposition: attachment; filename="vcard.vcf"

begin: vcard
fn: Gary Liechty
n: Liechty;Gary
org: Allied High Tech Products, Inc
adr: 2376 E. Pacifica Place;;PO Box 4608;Rancho Dominguez;CA;90220;USA
email;internet: garyliechty-at-worldnet.att.net
title: Product Application Specialist
tel;work: 800-675-1118
tel;fax: 310-762-6808
x-mozilla-cpt: ;0
x-mozilla-html: FALSE
version: 2.1
end: vcard


--------------350E4D479573B0548093D7E5--






From: Walck. Scott D. :      walck-at-ppg.com
Date: Fri, 4 Dec 1998 14:44:00 -0500
Subject: Low voltage Backscattered Imaging in FEG-SEMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am evaluating FEG semi-in-lens SEMs for possible purchase. I am
interested in BSE imaging at low acc. voltages ( {3kV). I would like to get
some feedback as to people's experiences with using BSE mode at low
voltages. Specifically, what kind of detectors are being used and what kind
of results? What is the lowest acc. voltage that you can reasonably use and
still get good results.
Thanks,





From: Gary Liechty :      garyliechty-at-worldnet.att.net
Date: Fri, 04 Dec 1998 13:06:42 -0800
Subject: Re: alloys

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------810728A47D701DBA5751A9F9
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

Hello Grace,

Contact ASM International in Ohio, 800-336-5152 for infomation on books that
will have what you are seeking.

Yours Truly,

Gary Liechty
Allied High Tech Products, Inc.

Grace Kennedy wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} This is a basic question for the materials scientists. I am interested in
} finding a simple reference which might delineate the different types of
} stainless steel, including relative hardness, tensile strength, etc. I'm
} also interested in seeing the same information plus the electrical
} characteristics of the various alloys of indium. In this case, melting
} points are of interest, and the exact formulae of "Wood's metal" and
} another that we use called "Cerralloy". Many thanks- Grace



--------------810728A47D701DBA5751A9F9
Content-Type: text/x-vcard; charset=us-ascii; name="vcard.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for Gary Liechty
Content-Disposition: attachment; filename="vcard.vcf"

begin: vcard
fn: Gary Liechty
n: Liechty;Gary
org: Allied High Tech Products, Inc
adr: 2376 E. Pacifica Place;;PO Box 4608;Rancho Dominguez;CA;90220;USA
email;internet: garyliechty-at-worldnet.att.net
title: Product Application Specialist
tel;work: 800-675-1118
tel;fax: 310-762-6808
x-mozilla-cpt: ;0
x-mozilla-html: FALSE
version: 2.1
end: vcard


--------------810728A47D701DBA5751A9F9--






From: Mark Darus :      DARUSM-at-cle.lg.bfg.com
Date: Fri, 04 Dec 1998 16:02:08 -0500
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Fri, 04 Dec 1998 13:18:46 -0800
Subject: Re: Query: How much better: WDS than EDS for SODIUM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Jacob,
The improvement of the WDX peak-to-background ratio over the EDX is even
more because the background is much lower. The detection limit is much
lower, so it is easier to quantify levels such as yours, which are close to
the EDX detection level. The problems of WDX are first: high cost and
second:, the need for a flat sample with exact geometry to the electron
beam. You must also run standards that are exactly the same geometry. The
WDX is much fussier than EDX on sample geometry and the analyses generally
take much longer, but have higher precision and lower detection level. I
have no experience with biological samples on the WDX, but there is a
listserver for microprobe analysis, i.e. WDX analysis. To subscribe to the
microprobe listserver use: microprobe-at-www.microanalysis.org, to post to it
use: microprobe-at-microanalysis.org. I don't know of any WDX spectrometers for
donation, there aren't that many around and they have to fit your SEM model.

You wrote:
}
} We are considering looking for a Wavelength-Dispersive Spectrometer to
} improve our detection of sodium in biological samples, reasoning that the
} better energy resolution of the detector will result in narrower, taller
} peaks with the same background level, yielding improved peak-to-background
} ratios.
}
} Our samples are frozen hydrated and dried biological tissues; we would like
} to be able to differentiate 50mMolar from 150mMolar sodium (0.25 and .75%
} by weight in hydrated samples).
}
} Our present EDS detector has a semi-thin window and is doing better than
} average for sodium and we would prefer not going to an ultrathin window or
} windowless detector.
}
} Might we expect a significant improvement with WDS?
}
} Does anyone know of a spectrometer that might be donated?
}
} Your comments and suggestions will be appreciated.
}
} Thanks,
}
} Jacob
}
} Jacob Bastacky, M.D.

Regards,
Mary
Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
fax: 604-822-3619
e-mail: mager-at-interchg.ubc.ca






From: Barbara Plowman :      Bplowman-at-unix1.cc.uop.edu
Date: Fri, 04 Dec 1998 13:44:42 -0800
Subject: Immuno

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Listserver!
I am doing some immunohistochemistry and have made two attempts but with no
luck at getting any gold label. My question is whether to use a lower
dilution gold-antibody or extend the incubation from 4 hours to overnight or
both? I am a self-confessed rookie at this, so any information on the subject
would be appreciated .





From: Lucille A. Giannuzzi :      lag-at-mail.ucf.edu
Date: Fri, 4 Dec 1998 15:51:29 -0500
Subject: Position Available: University of Central Florida

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Position Available
University of Central Florida
Advanced Materials Processing and Analysis Center
(AMPAC)

The Advanced Materials Processing and Analysis Center (AMPAC) is seeking
candidates for a position as Assistant in Research to support the Materials
Characterization Facility (MCF). The individual must have a Bachelor
degree from an accredited institution in an appropriate area of study
related to surface science and materials characterization, and have at
least three years of experience in materials characterization or related
areas. The person will be responsible for establishing and maintaining
laboratory infrastructure, laboratory maintenance, equipment installation,
operation and maintenance, and be an investigator in contracted research.
The person must be able to conduct independent research and development in
materials characterization with an emphasis in electron beam technology and
have the ability to instruct students in the use and interpretation of
data. The individual must have extensive knowledge and experience in the
operation, maintenance and interpretation of materials characterization
equipment. The salary range starts at $35,000 and is negotiable dependent
on level and experience of the individual. The application opening date is
December 18, 1998 and the closing date is December 31, 1998. Applicants
should send a vitae and references to Dr. Lucille Giannuzzi, Director,
Materials Characterization Facility, AMPAC Administrative Offices, 12424
Research Parkway, Suite 408, Orlando, FL 32826. UCF is an equal
opportunity/affirmative action employer. As an agency of the State of
Florida, all application materials and selection procedures are available
for public review.


**************************************************************************
Lucille A. Giannuzzi, Ph.D. phone: 407 823-4770
University of Central Florida fax: 407 823-0208
Materials Science Program email: lag-at-pegasus.cc.ucf.edu
Dept. of Mechanical and Aerospace Eng.
Orlando, FL 32816-2450
**************************************************************************







From: kennedy-at-nsi.edu (Grace Kennedy)
Date: Fri, 4 Dec 1998 14:40:57 -0800
Subject: alloy replies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks to all of you who kindly replied to my question, I appreciate the help.







From: shAf :      mshaf-at-darkwing.uoregon.edu
Date: Fri, 4 Dec 1998 14:37:12 -0800
Subject: RE: Low voltage Backscattered Imaging in FEG-SEMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Scott asks ...
}
} ...
}
} I am evaluating FEG semi-in-lens SEMs for possible purchase. I am
} interested in BSE imaging at low acc. voltages ( {3kV). ...

I am unfamiliar with the geometry for "semi-in-lens" vs a BSE
detector which would need line-of-sight to the interaction spot on
the sample. How does this work???

cheerios, shAf

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - ICQ 210524
Geological Science's Electron Probe Facility - University of Oregon
mshaf-at-darkwing.uoregon.edu - http://darkwing.uoregon.edu/~mshaf/







From: Elaine Humphrey :      ech-at-unixg.ubc.ca
Date: Fri, 4 Dec 1998 15:31:30 -0800
Subject: grids dull or shiny?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all
We have recently had a number of users who have come from different places
in the world, and use different conventions for whether sections/formar go
on the shiny side of the grid or the dull side.

This lab uses the dull side, and has considered that the convention. If it
ain't broke don't fix it.

As long as the user is consistent with his/her own grids, there is no
problem, but when looking at collegues grids from another lab, it can lead
to a few annoyances.

We have been making our own formvar or pioloform-coated grids and putting
the formvar or pioloform on the dull side. One of my users has bought
made-up-grids where the coating was on the shiny side. At least two other
labs here use the shiny side for sections.

Is there a difference to the sides of the grid that makes one side better
than the other?
Elaine


Dr. Elaine Humphrey
Biosciences Electron Microscopy Facility
University of British Columbia
6270 University Blvd, mail-stop Botany
Vancouver, BC
CANADA, V6T 1Z4
Phone: 604-822-3354
FAX: 604-822-6089
e-mail: ech-at-unixg.ubc.ca







From: bozzola-at-siu.edu (John J. Bozzola)
Date: Fri, 4 Dec 1998 20:06:08 -0600
Subject: Re: grids dull or shiny?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dr. Elaine Humphrey asked:

We have recently had a number of users who have come from different places
in the world, and use different conventions for whether sections/formar go
on the shiny side of the grid or the dull side.

RESPONSE:

This is a good basic question - one that many microscopists have asked when
coating grids with various plastic substrates. This question has even been
studied in a scientific manner over the years and the conclusion has
generally been that the dull side appears to hold on to the film more
tenaceously than the shiny side. Now, I must admit that I started out using
the shiny side and resisted going dull for many years (in reference to
grids not sense of humor). I finally did a test and found that there was
less drift and better film adhesion on the dull side.

Now a caveat: examine carefully the two sides of the grids and check if
there are noticeable burrs or curvature to the grids (especially the
latter). You should place the plastic film so that it is in maximal contact
with the grid. So, if the grid is slightly bent, make sure the film goes on
the convex side. Reason: on the concave side the film is suspended over the
concavity and you will get a lot of drift. I have noticed that some slot
grids have such a concavity and it may require you to place the film on the
"shiny" side in this case.

So, generally speaking, place the film on the dull side of the grid unless
the grid is dished or bent. In order to further facilitate adhesion of the
film, some people place a drop of diluted plastic (like Formvar at 0.02%)
on top of grids while on a filter paper. When dried, the thin plastic film
has an affinity for the substrate or film that you finally place onto the
grid for supporting sections, etc.




####################################################################
John J. Bozzola, Ph.D., Director
Center for Electron Microscopy
Neckers Building, Room 146 - B Wing
Southern Illinois University
Carbondale, IL 62901
U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################







From: Z.L. Wang :      zhong.wang-at-mse.gatech.edu
Date: Fri, 04 Dec 1998 23:09:44 -0800
Subject: Structural evolution in Smart Materials - a new book

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


{bold} {italic} FUNCTIONAL AND SMART MATERIALS

-Structural evolution and structure analysis

by Z.L. Wang and Z.C. Kang

{/italic} {/bold} (Plenum Press, 1998)



ISMN 0-306-45651-6, pp. 514

Ordering information: http://www.plenum.com


{bold} {italic} Physics Today {/italic} {/bold} (Nov. 1998, p. 70):

"... this book is a unique, cutting-edge text on smart materials ... it
is recommended as an adjunct to device design books used for engineers as
well as scientists during the development of smart devices and
structures."


{bold} {italic} {bigger} Science {/bigger} {/italic} {/bold} (Vol. 281, 10 July
1998, p. 181):

"The authors consider the atomic scale crystal structure and chemistry of
oxides with physical and chemical properties that are sensitive to
changes in the environment such as temperature, pressure, electric or
magnetic fields, pH, and optical wavelength. They explain relationships
among different structures and explore approaches to characterizing and
synthesizing these important components for electronic devices"


{bold} {italic} COMMENTS FROM EXPERTS IN THE FIELD:

{/italic} {/bold}

"Both new in concept and timely in publication....Bring together, for the
first time, the fundamental of atomic scale crystal structure and
chemistry... A cutting-edge text at the forefront of the modern materials
revolution"


- Professor David B. Williams, Lehigh University, USA


"Unique...focuses specifically on the intrinsic connections among several
crystal structure systems and their evolution behavior...Fills a gap left
in the field ... This book will be a basic reference in the domain of
oxides which are to be the basis of functional and smart materials"


- Professor C. Boulesteix, Universite Aix-Marseille, France.


"In materials science the spotlight is on functional and smart materials,
since they are important components for electronic devices. The textbook
by Wang and Kang summarizes all types of known functional materials and
describes the structure evolution problems. A large section of the book
is devoted to structural characterization focusing on transmission
electron microscopy, the main field of expertise of the author. The book
is extremely valuable for materials scientists working on functional
oxide materials, studying the structure, structure evolution and defects.
It may serve also as an interesting textbook for teaching since it gives
a good overview of this field which is of increasing importance. The
clarity of its writing style should make it ideally suited for graduate
students."


- Professor Manfred Ruhle, Institte of Werkstoffwissenschaft, Germany=20



Contents

Part I

STRUCTURE AND STRUCTURAL EVOLUTION=20


1. Structure, Bonding and Properties

1.1 Crystal structure

1.2 Structure and chemical composition

1.2.1 Stoichiometric phases

1.2.2 Non-stoichiometric phases

1.3 Coordination number and coordination polyhedron

1.4 Isotypism and polymorphism

1.5 Structure and chemical bonding

1.5.1 Bonding and ion radius

1.5.2 Lattice energy of an ionic compound

1.5.3 Geometric consideration of a structure

1.5.4 The rules of Pauling and Baur

1.5.5 Covalent bonding

1.6 Ligand field theory

1.6.1 Octahedral coordination

1.6.2 Tetrahedral coordination

1.6.3 Square coordination

1.7 Ligand field stabilization energy

1.8 Coordination polyhedra of transition metals

1.9 Molecular orbital theory of bonding

1.9.1 Molecular orbitals

1.9.2 Hybridization

1.10 Band theory

1.10.1 The Peierls distortion

1.10.2 Two- and three-dimensional bonds

1.11 Mixed valent compounds and functional materials

1.12 Structure transformation and stability

1.12.1 Phase diagram

1.12.2 Thermodynamic stability

1.13 Properties of materials

1.13.1 Mechanical Property

1.13.2 Magnetic property

1.13.3 Piezoelectric property

1.13.4 Ferroelectric property

1.13.5 Optical property

1.13.6 Electric property

1.14 Structure and property

1.15 Functional materials

1.15.1 Characteristics of functional materials

1.15.2 Structural evolution and functionality

1.16 Summary


2. Sodium Chloride and Rutile Related Structure Systems

2.1 Rock salt structure

2.2 Non-stoichiometric compounds with sodium chloride structure

2.3 Rutile structure and its family

2.4 Characteristics of rutile structures

2.5 Evolution of rutile-type structures

2.6 Non-stoichiometry and crystallographic shear planes

2.7 Summary


3. Perovskite and Related Structure Systems=20

3.1 Characteristics of ABO3 type perovskite structure

3.1.1 Vertex-sharing of oxygen octahedra

3.1.2 Unit cell by taking A cation as the origin

3.1.3 Oxygen cubic close-packing

3.1.4 Anion close-packing and formation of tetrahedron and octahedron

3.2 Possible types of anion-deficient perovskite structures

3.2.1 The 14 fundamental structure units

3.2.2 Constructing the family of perovskite related structures

3.3 The tolerance factor

3.4 Functional materials with perovskite-like structures

3.4.1 Ferroelectricity and ferroelectric compounds

3.4.2 Ferromagnetism and ferromagnetic compounds

3.4.3 Insulator to conductor transition

3.4.4 Conductive perovskites

3.4.4.1 Valence disproportionationality

3.4.4.2 Dimensionality

3.4.4.3 Building the structures of high temperature superconductors using
perovskite structure units

3.4.5 Magnetostrictive, electrostrictive and piezoelectric actuator
materials

3.4.6 Optically switchable compounds

3.5 Doping and oxygen vacancies

3.6 Giant magnetoresistance (GMR) and colossal magnetoresistance
(CMR)

3.7 Oxygen migration and ionic conductivity of perovskites=20

3.8 Anion deficiency induced perovskite to brownmillerite structural
evolution=20

3.9 Ordered structural evolution introduced by cation substitution

3.10 Sodium chloride, rutile and perovskite structures

3.10.1 Linkage and comparison

3.10.2 Constructing new materials by tailoring

3.11 Summary


4. Fluorite-Type and Related Structure Systems=20

4.1. Basic fluorite structure

4.2 Fluorite structure with anion deficiency

4.2.1 Oxygen migration in fluorite structure

4.2.2 Modules of fluorite structure with oxygen deficiency

4.2.3 Pyrochlores and C-type rare earth sesquioxide structures

4.3 Characteristics of fluorite and fluorite-related structures

4.3.1 Thermodynamic property

4.3.2 Surface character of rare earth oxides

4.3.3 Disproportionation of rare earth high oxides

4.3.4 Switchable chemical reaction as an oxygen pump

4.4 Structural and compositional principles of rare earth homologous
higher oxides

4.4.1 Compositional principle of the homologoues phases

4.4.2 The modular juxtaposition rules

4.4.3 Building supercell structure using modules

4.5 Applications of the juxtaposition rules to known structures

4.5.1 R7O12 phase with n =3D 7 and m =3D 1

4.5.2 R9O16 phase with n =3D 9 and m =3D 1

4.5.3 R11O20 phase with n =3D 11 and m =3D 1

4.5.4 R40O72 phase with n =3D 40 and m =3D 4

4.5.5 R24O44 phase with n =3D 24 and m =3D 2

4.6 Predicting undetermined structures using the proposed modules

4.6.1 b-polymorph with m =3D 4

4.6.2 Undetermined structure with n =3D 19

4.6.3 Undetermined structure with n =3D 16

4.6.4 Undetermined structure with n =3D 62 and m =3D 6

4.6.5 Non-stoichiometric a-phase

4.7 Ternary mixed rare earth oxides

4.7.1 Rare earth mixed ternary oxides and oxygen storage

4.7.2 Cation coordination number and arrangements of modules

4.8 Perovskite, fluorite structures and spinel structures

4.8.1 Structure comparison

4.8.2 Superexchange interaction and magnetism

4.9 Summary


5. From Structural Units to Materials Engineering via Soft-Chemistry

5.1 Principle of soft chemistry

5.2 Sol-gel process

5.3 Colloidal route for preparation of monodispersive spherical
particles

5.4 Intercalation and pillaring processes

5.5 Self-assembled nanocrystal engineered superlattice thin films

5.5.1 Passivated metal nanocrystals

5.5.2 Passivated semiconductors nanocrystals

5.5.3 Passivated magnetic nanocrystals

5.5.4 Magnetic Co particles

5.5.5 Magnetic iron oxides

5.6 Preparation of nanoparticles by aerosol technique

5.7 Summary

Part II

STRUCTURE CHARACTERIZATIONS


6. Electron Crystallography for Structure Analysis=20

6.1 Electron diffraction in structure analysis

6.1.1 Single scattering theory

6.1.2 Reciprocal space

6.1.3 Bragg's law and Ewald sphere

6.1.4 Indexing electron diffraction patterns

6.1.5 Diffraction from twinned crystals

6.2 Diffraction contrast and defect analysis

6.2.1 Defects and dislocations in materials

6.2.2 Diffraction contrast

6.2.3 Two-beam condition for imaging defects and dislocations

6.2.4 Determination of Burgers vector

6.2.5 Weak beam imaging

6.3 Atomic-resolution structure imaging and structure analysis

6.3.1 Phase contrast

6.3.2 Abbe's imaging theory

6.3.3 Aberration and information transfer in TEM

6.3.4 Contrast transfer function and image resolution

6.3.5 Envelope function and information transfer

6.3.6 Source coherence in lattice imaging

6.3.7 Projected charge density approximation

6.3.8 Multislice theory for transmission electron imaging

6.3.9 Image simulation and structure determination

6.3.10 Image calculation of imperfect crystal and interface

6.3.11 Energy-filtered electron lattice imaging

6.3.12 Limitation of HRTEM

6.4 Electron holography

6.4.1 Principle of off-axis holography in TEM

6.4.2 Improvement of image resolution

6.4.3 Imaging electrostatic field and charge distribution

6.4.4 Imaging spontaneous polarization at domain boundaries in
ferroelectrics

6.4.5 Imaging magnetic domains and flux lines

6.5 Convergent beam electron microdiffraction

6.5.1 Symmetry analysis

6.5.2 Measurement of lattice parameters

6.5.3 Bloch wave theory and quantitative CBED

6.5.4 Solid state bonding and charge redistribution

6.5.5 Determination of Burgers vector

6.5.6 Measurement of specimen thickness

6.6 Summary


7. Structure analysis of functional materials=20

7.1 Interface and grain boundary analysis

7.1.1 Kikuchi pattern and grain boundary analysis

7.1.2 General description of a grain boundary

7.1.3 The O-lattice theory

7.1.4 Coincidence-site lattice theory

7.2 Modulation and domain structure

7.2.1 Structural modulation

7.2.2 Domains formed by anisotropic crystal structure

7.2.3 Boundaries of structure domains

7.3 Superstructure and long-range ordering

7.3.1 3-D superstructure analysis by a double-pattern technique

7.3.2 3-D superstructure analysis by a single-pattern technique

7.3.4 Long-range ordering of cation substitutions

7.4 Oxygen vacancies and short-range ordering

7.4.1 Kinematical diffraction theory of diffuse scattering

7.4.2 Geometrical description of diffuse scattering

7.4.3 Calculation of short-range ordering parameter

7.4.4 HRTEM study of short-range order

7.5 Effects of substrate on thin film growth

7.5.1 Lattice mismatch and interface dislocations

7.5.2 Nucleation and growth of defects from substrate surfaces

7.5.3 Linkage of domain boundaries with interface dislocations

7.5.4 Linkage of interface dislocations with planar defects

7.6 In-situ observation of structure evolution

7.6.1 Temperature driven structure transformation

7.6.2 Electric field driven structure transformation

7.6.3 Magnetic moment of the specimen

7.7 Failure analysis of devices

7.8 Imaging surfaces of oxides

7.9 Summary


8. Chemical and Valence Structure Analysis of Functional Materials

8.1 Inelastic excitation processes in electron scattering

8.2 Energy dispersive x-ray microanalysis (EDS)

8.2.1 Composition analysis

8.2.2 Atom location by channeling enhanced microanalysis (ALCHEMI)

8.3 Valence excitation EELS

8.3.1 Classical electron energy-loss theory

8.3.2 Surface plasmon excitation

8.3.3 Measurement of dielectric function

8.4 Atomic inner shell excitation in EELS

8.4.1 Composition analysis

8.4.2 Near edge fine structure and bonding in crystals

8.5 Quantitative determination of valences in a mixed valent=20
compound

8.5.1 White lines of transition metals

8.5.2 The occupation number of the d-band electrons

8.5.3 White line intensity and intrinsic magnetic moment

8.5.4 Double derivative spectrum for calculation of white line=20
intensity

8.6 Nano-probe analysis of interfaces and grain boundaries

8.7 Chemical sensitive imaging in STEM

8.8 Energy filtered electron imaging in TEM

8.8.1 Composition-sensitive imaging using valence-loss electrons

8.8.2 Composition-sensitive imaging using inner-shell ionization edge
electrons

8.8.3 Mapping of bonding and valence state

8.9 Phonon scattering and chemical sensitive imaging

8.9.1 'Z-contrast' imaging in STEM

8.9.2 High-angle dark-field conical scan imaging in TEM

8.10 Conjunction use of various techniques for structure refinement of
La8Sr8Co16O36 - an example

8.11 Summary

=09

APPENDIXES

A: Physical constants, electron wavelengths and wave numbers

B1: Crystallographic structure systems

B2: FORTRAN program for calculating crystallographic data

C: Electron diffraction patterns for several types of crystal=20
structures

D: FORTRAN program for calculating single valence-loss EELS spectra in
TEM


Introduction


Smart systems and smart materials

Smart structures are a new emerging materials system which combines
contemporary materials science with information science. The smart=20
system is composed of sensing, processing, actuating, feedback,
self-diagnosing and self-recovering subsystems. The system uses the
functional properties of advanced materials to achieve high performances
with capabilities of recognition, discrimination, and adjustification in
response to a change of its environment. Each component of this system
must have functionality, and the entire system is integrated to perform a
self-controlled smart action, similar to a living creature who can
"think", make judgment and take actions. A smart system can be considered
as a design philosophy that emphasizes predictivity, adaptivity and
repetivity.=20

A smart system/structure is defined to be a non-biological physical
structure having the following attributes: (1) a definite purpose; (ii)
means and imperative to achieve that purpose; and (iii) a biological
pattern of functioning (Spillman et al., 1996). Smart materials are a
subset of the smart system, i.e. smart structures at the microscopic or
mesoscopic scales. Smart system is a non-biological structure which means
that the system functions as a biological system rather than the pattern
of functioning of a Turning machine. A smart material is a physical
structure having (i) a definite purpose, (ii) means of imperative to
achieve that purpose, and (iii) the pattern of functioning of a universal
computer or Turning machine. Such a material will generally include at
least one structural element, some means of sensing the environment
and/or its own state, and some type of processing and adaptive control
algorithm.=20

Science and technology in the 21st century will rely heavily on the
development of new materials that are expected to respond to the
environmental changes and manifest their own functions according to the
optimum conditions. The development of smart materials will undoubtedly
be an essential task in many fields of science and technology such as
information science, microelectronics, computer science, medical
treatment, life science, energy, transportation, safety engineering and
military technologies. Materials development in the future, therefore,
should be directed toward creation of hyperfunctional materials which
surpass even biological organ in some aspects. The current materials
research is to develop various pathways that will lead the modern
technology toward the smart system.=20


Functional materials

Functional materials are distinctly different from structural materials,
and their physical and chemical properties are sensitive to a change in
the environment such as temperature, pressure, electric field, magnetic
field, optical wavelength, adsorbed gas molecules and the pH value. The
functional materials utilize the native properties and functions of their
own to achieve an intelligent action. Functional materials cover a
broader range of materlas than the smart materials illustrated above.
Besides the materials belogning to the smart structure, any materials
that have functionality are attributed to functional materials, such as
the ferroelectricic BaTiO3, the magnetic field sensor of La1-xCaxMnO3,
surface acoustic wave sensor of LiNbO3, liquid petroleum gas sensor of
Pd-dopped SnO2, semiconductor light detectors (CdS, CdTe), high
temperature piezoelectric Ta2O5, fast-ion conductor Y2(SnyTi1-y)2O7
(pyrochlore structure), the electric voltage induced reversible colouring
of WO3, and high temperature superconductors etc. Functional materials
cover a wide range of organic and inorganic materials. This book focuses
only on oxide functional materials.

In recent years, techniques for epitaxial crystal growth have made it
possible to grow oxides and metal thin films on silicon substrates, and
this is the first step to integrate functional materials with the logic
system. Preparations of complex oxides with functionality are a key
challenge for materials development. Searching new routes to prepare
materials and understanding the relationship between the structures and
the properties are equally important. A key requirement in preparations
of materials is to control the structural and compositional evolution for
achieving superior properties. "Soft chemistry" has shown a great success
in fabricating functional and nanophase materials. Nanocrystal engineered
materials are a new trend of materials research, aiming to improve the
performances of materials by several orders of magnitudes.=20


Mixed valences and functionality

Crystal structure usually refers to two aspects of information, one
is the symmetry and distribution of atoms in the unit cell and the other
is the bonding between atoms. Thermordynalically, the enthalpy of the
system is determined by the bonding between atoms, while the entropy is
determined by the atomic lattice configurations of the crystal.
Thermodynamic rules select the possible stable phases, and the phase
stability is strongly affected by bonding. A single element has a certain
electron configuration. When several different elements form a molecule,
the electronic structure of this cluster is very different from any of
its original elemental configuration because of the transfer and/or
sharing of valence electrons among atoms. In general, only the valence
electrons are most critical to bonding, the distribution and motion of
valence electrons are usually described by the molecular orbitals. These
valence states and molecular orbits are responsible for the functional
properties of the molecule. The ligand field theory is designed to
describe the molecular structure of an atom cluster.=20

When different elements are combined to form a crystalline solid in
which the atoms or atom groups (or molecules) are bonded together to form
a three-dimensional (3-D) structure with specified symmetries, the
properties of the solid would depend on both the electronic structure of
the atoms or atom groups and their spatial distribution. The molecular
orbital theory and band structure theory are usually applied to elucidate
the relationship between the structure and the properties. Based on the
electron band structure, inorganic materials can be classified as
conductors, semiconductors and insulators. If a change is made in the
crystal structure so that the band gap is reduced or eliminated, a
transition from an insulator to a conductor is possible. Modifying a
crystal structure can be performed by changing either the spatial
distribution of atoms (such as bonding angles, bonding lengths and
symmetry of atom arrangement) or chemical composition (such as from
stoichiometry to nonstoichiometry). All these changes are referred to as
structural evolution, which is closely related to the properties of the
materials.=20

Many functional properties of inorganic materials are determined by the
elements with mixed valences in the structure unit, by which we mean
that an element has two or more different valences while forming a
compound. Valence mixture refers to a case in which several elements
have different valences but each one only has a single valence. In the
periodic table of fundamental elements, 40 elements can form mixed valent
compounds, transition d-block elements and lanthanide (Eu, Yb, Ce, Pr, Tb
etc.) are typical examples. Modern inorganic chemistry has shown that the
oxidation state of any element can be modified under special conditions.
Many oxide functional materials contain elements with mixed valences.
This is a typical difference between the functional materials and the
structural materials. This book is about the structural evolution of
compounds containing mixed valent elements, such as transition and rare
earth elements. =20

The concept of mixed valent chemistry offers a pathway to design and
synthesize new compounds with unique optical, electric, or magnetic
properties. Research in functional materials in its broad sense always
depends on the conception and synthesis of interesting novel mixed-valent
compounds. The discovery of high temperature superconductor compounds is
a fascinating successful example of the mixed valence chemistry. We
believe that exploring the possible structures of mixed valent compounds
and their evolution behaviors may lead to many pathways to synthesis new
functional materials.=20


The scope of the book

Searching for new functional materials is a challenge for the
development of smart systems. To guide this searching, a clear
understanding about the relationship between the physical properties and
the atomic-scale structure of the materials is desperately needed. This
book is about the intrinsic connections among several crystal structure
systems commonly used in functional materials and their evolution
behaviors. The book is not intended as a source for listing all the
existing functional materials, instead its goal is to reveal the
principles for engineering and controlling functional materials from the
fundamental structure units. The functional materials are described from
the mixed valences and stoichiometry points of view to understand the
structural evolution and transformation of different materials systems.
The mixed valent compounds are elucidated as the fundamentals for
performing unique functionality.=20

We have written this book with a strong conviction that functional
materials system is a future direction of the multidisciplinary research
involving physics, chemistry, materials science, electrical engineering
and biological science, with an emphasis on molecular and unit cell
designs. There are numerous books describing the properties,
preparations, electronic structures and crystal structures of transition
and rare earth metals and their oxides. To expert in the field, this book
is written by addressing the issues that have not been described
systematically in existing books. Our aim is to explain the intrinsic
connections among different structures. The goal is to explore new routes
for synthesizing functional materials from the fundamental structure
building blocks (or modules). The book aims to illustrate not only the
role played by crystal structure in property control of functional
materials, but also the structure determination using advanced
transmission electron microscopy and spectroscopy techniques. The latter
also has critical importance for device failure analysis in modern
industry.

Accordingly, the book is written into two parts. Part I concentrates on
the structure and structural evolution of oxides functional materials
belonging to NaCl-, rutile-, perovskite and CaF2-type structures and the
related. Although our analysis is focused on the structure systems
outlined above, spinel and corundum structures are also briefly described
in these chapters except the wurtzite structure (BeO and ZnO) because of
its limited evolution characteristics. Part II is on crystallographic and
chemical structure characterizations of oxide functional materials, which
are needed to understand the experimental approaches for exploring the
structure evolution. Both parts are written to ensure the coherency of
the whole book.

Part I is composed of five chapters. In Chapter 1, fundamental concepts
are introduced on crystallographic structures, chemical structures,
bonding, molecular orbital and ligand field, mixed valences, materials
properties and the fundamental characteristics of functional materials.
This chapter is designed as the preliminary preparation for the
discussions to be outlined in future chapters. The characteristics of
functional materials are given for distinguish them from structural
materials. Chapter 2 starts with the simplest oxides with NaCl structure
for illustrating the stacking of cations and anions in constructing the
unit cell. Then, rutile related structures are introduced with an
emphasis on their structural evolution. Chapters 3 is on perovskite and
related structures. The characteristics of the ABO3 perovskites are
thoroughly analyzed to reveal the intrinsic connections among the A, B
and O elements and the roles played by the octahedron in constructing the
unit cells. The alternative stacking of the close-packed (AO3)4- and B
layers is analyzed in detail, and a total of 14 fundamental structure
units are extracted with introducing of anion deficiency, which are the
building blocks for constructing the unit cells of complex functional
materials. The tailoring of cation octahedra can give the structure of a
variety of compounds. The typical properties of perovskites are also
described and their relations with the crystal structure are elucidated.
This chapter is the basis for understanding the mechanisms that control
cation substitution, creation of oxygen vacancies and mixed valences in
the perovskite family.

Similarly, the fluorite related structures are analyzed in Chapter 4,
and a total of 12 fluorite modules (or units) with anion deficiencies are
proposed, which together with the perfect fluorite are the 13 basic
building blocks for constructing the structures of rare earth homologoues
higher oxides. Following the structural and compositional principles
outlined by Kang and Eyring (1995, 1996), geometrical assembling of these
modules can reproduce not only the known crystal structures of
homologoues phases RnO2n-2m but, more importantly, predict the structures
of some phases whose structures are unknown. This chapter proves that a
complex superstructure can be disassembled into some fundamental modules,
which can be derived at the first place from the basic fluorite
structure. Therefore, by revealing the intrinsic connection among the
same class of phases one may be able to predict and design new
structures. The surface property and oxygen migration can be better
understood from the view of module combination. At the end of Chapter 4,
the structures of perovskite, fluorite and spinel are compared with each
other and a brief introduction about spinel is given.=20

Chapter 5 focuses on the introduction of soft-chemistry (sol-gel,
pillaring and grafting, intercalation and deintercalation) for
synthesizing and designing new materials based on the fundamental
structure modules. The chapter aims to using the understanding about
structural evolution described in previous chapters to develop new
materials systems that are expected to have functionality. Preparation of
nanophase materials is also described with an emphasis on self-assembled
nanocrystal superlattices (or arrays). This is a new trend of materials
research on nanocrystal engineered (or patterned) materials.

Part II is composed of Chapters 6-8. Chapter 6 aims to introduce the
fundamentals of structure analysis using transmission electron microscopy
and associated techniques, including electron diffraction, diffraction
contrast for defect analysis, atomic resolution lattice imaging for
interface studies, electron holography for studying ferroelectric and
ferromagnetic materials, and convergent beam electron diffraction
technique for mapping charge redistribution and bonding in crystalline
materials. While introducing these fundamental techniques, our focus is
on the applications of these techniques for the analysis of functional
materials. This chapter also serves as an educational chapter to the
readers who have different background in solid state chemistry, materials
processing and structure analysis, for correctly using the available
tools for solving the problems in oxide functional materials. The
applications of the techniques introduced in Chapter 6 are described in
Chapter 7 case-by-case to cover the structural characteristics of oxide
functional materials, including grain boundaries, interfaces, domains,
structure transformations and surfaces. Analysis of long-range
superstructures and short-range order of point defects is the emphasis of
this chapter because they are closely related to the structural evolution
described in Chapters 2-4. Numerous examples are shown to illustrate the
techniques for solving the structure problems belonging to functional
materials.

Chapter 8 focuses on the chemical and bonding analysis of mixed valent
oxides at a high spatial resolution using the energy dispersive x-ray
spectroscopy (EDS) and electron energy-loss spectroscopy (EELS). The near
edge fine structure observed in EELS is a sensitive technique for
detecting the valence band structure from a small region, and it also
allows the analysis of interface electronic structures. The observed
white lines in EELS can be used to determine the occupations of the 3d
and 4d orbitals (e.g., valence states) of transition and rare earth
elements. This information together with the imaging and diffraction data
from HRTEM can be applied to determine the ordered structure induced by
ordered anion vacancies. Finally, it is emphasized that the structure
information provided by imaging and diffraction techniques (Chapters 6
and 7) must be integrated with the chemical information provided by EDS
and EELS. The structure refinement of an anion deficient phase
La8Sr8Co16O36 is demonstrated as an example. A combination of TEM results
with x-ray diffraction or neutron diffraction data and theoretical
modeling as well is likely to give a unique and reliable solution to a
material's problem. It is concluded that the structural evolution in a
complex system is intrinsically dominated by the combinations of the
fundamental structural modules. This is the focus of this book.








From: Webster :      pwebster-at-mailhouse.hei.org
Date: 05 Dec 98 00:47:56 -0800
Subject: RE: grids dull or shiny?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
MSA listserver submission {Microscopy-at-Sparc5.Microscopy.Com}
X-Mailer: QuickMail Pro 1.5.3 (Mac)
X-Priority: 3
MIME-Version: 1.0
Reply-To: Webster {pwebster-at-mailhouse.hei.org}
Content-Transfer-Encoding: quoted-printable
Content-Type: text/plain; charset="US-Ascii"



From: Webster :      pwebster-at-mailhouse.hei.org
Date: 05 Dec 98 00:47:56 -0800
Subject: RE: grids dull or shiny?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Reply to: RE: grids dull or shiny?





Elaine Humphrey wrote:
} Hello all
} We have recently had a number of users who have come from different =
places
} in the world, and use different conventions for whether sections/formar =
go
} on the shiny side of the grid or the dull side.
}
} This lab uses the dull side, and has considered that the convention. If =
it
} ain't broke don't fix it.
}
} As long as the user is consistent with his/her own grids, there is no
} problem, but when looking at collegues grids from another lab, it can =
lead
} to a few annoyances.
}
} We have been making our own formvar or pioloform-coated grids and putting
} the formvar or pioloform on the dull side. One of my users has bought
} made-up-grids where the coating was on the shiny side. At least two =
other
} labs here use the shiny side for sections.
}
} Is there a difference to the sides of the grid that makes one side better
} than the other?
} Elaine
}
}
} Dr. Elaine Humphrey
} Biosciences Electron Microscopy Facility
} University of British Columbia
} 6270 University Blvd, mail-stop Botany
} Vancouver, BC
} CANADA, V6T 1Z4
} Phone: 604-822-3354
} FAX: 604-822-6089
} e-mail: ech-at-unixg.ubc.ca
}
Reply;

Ter trivial answer to this question is that all experts are correct. =
Therefore either side is good. However, there is one instance where I =
have found this to be incorrect. Single slot grids have a polarity to =
them. The dull side has rounded edges wherease the shiny side is =
completely flat. Check it out. the immediate concequence is that if =
formvar is applied to the dull side it will eventually sag into the hole =
and stick to the support you store it on. On the shiny, flat side, this =
is not a problem. This might be true for the mesh grids too. I can't say =
more on this because I never looked. =

For the record, for routine use I usually put the film on the dull side =
too and have never had a problem. I know a colleague who swears that using =
the shiny side is best because it improves resolution! I do know that if =
the film has been damaged in some way then buffers will react with the =
metal. Hence the dogma of not using copper grids for immunocytochemistry. =
=

My advice is to believe all the experts and do what ever you want. Even =
better, become an expert and make everyone in your lab do it your way!
}
}
}
} RFC822 header
} -----------------------------------
}
} Received: from Sparc5.Microscopy.Com [206.69.208.10] by mailhouse.hei.=
org
} (SMTPD32-4.07) id A8EC245B023C; Fri, 04 Dec 1998 21:47:24 PST
} Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.=
11) =
} id RAA29721 for dist-Microscopy; Fri, 4 Dec 1998 17:19:29 -0600
} Received: from no_more_spam.com (Sparc5 [206.69.208.10]) by =
} Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id RAA29715 for =
} "MicroscopyFilteredEmail-at-msa.microscopy.com"; Fri, 4 Dec 1998 17:18:58 -=
0600
} Received: from mail.unixg.ubc.ca (mail.unixg.ubc.ca [137.82.27.14]) by =
} Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id RAA29707 for =
} {microscopy-at-sparc5.microscopy.com} ; Fri, 4 Dec 1998 17:18:44 -0600
} Received: from cbig-se.botany.ubc.ca ([137.82.136.14])
} by mail.unixg.ubc.ca with esmtp (Exim 1.92 #1)
} for microscopy-at-Sparc5.Microscopy.Com
} id 0zm4h8-0000tc-00; Fri, 4 Dec 1998 15:31:30 -0800
} X-Sender: ech-at-pop.unixg.ubc.ca
} Message-Id: {v0300780be8f1d9725816-at-[137.82.136.14]}
} Mime-Version: 1.0
} Content-Type: text/plain; charset=3D"us-ascii"
} To: microscopy-at-Sparc5.Microscopy.Com
} From: Elaine Humphrey {ech-at-unixg.ubc.ca}
} Subject: grids dull or shiny?
} Date: Fri, 4 Dec 1998 15:31:30 -0800
} Errors-to: Microscopy-request-at-sparc5.microscopy.com
} X-UIDL: 907881843
} Status: U
} =

Paul Webster, Ph.D
House Ear Institute
2100 West Third Street
Los Angeles, CA 90057
phone:213 273 8026
fax: 213 413 6739
e-mail: pwebster-at-hei.org






From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Sat, 5 Dec 1998 07:03:48 -0500
Subject: RE: Low voltage Backscattered Imaging in FEG-SEMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by arl-img-12.compuserve.com (8.8.6/8.8.6/2.16) id HAA12219;
Sat, 5 Dec 1998 07:04:07 -0500 (EST)


Hi Guys

Maybe I can help?

Semi in lens imaging, or dual detector imaging, relies upon the SE being
attracted to an detector which is situated above the final lens pole piec=
e.
Initially commercial instruments with two ET detectors used the field fr=
om
the final lens to "pull" SE into the upper detector. ISI were first with=

their SS Series and then Hitachi with the S570 took this route. In these=

instruments they allowed the specimen to be anywhere between 100% out of
lens through to what we would consider to be a WD of -5mm. That is 5mm u=
p
inside the pole piece! Problems arose if you were using magnetic
materials, hence the development of a twin detector system by Hitachi whi=
ch
used a field coil to drag electrons up to the detector and a pole piece
design which retained the lens field within the pole piece. =


The twin detector system gives the microscopist the best of all worlds.

1. The upper detector provides an opportunity to sift out the BSE an=
d
obtain a pretty pure SE image. The high lens strength at very short WD
(~3mm) enables the instruments to reach very high resolution levels. The
down side of this is, as SE are effected by charge, this mode is prone to=

charge problems.

2. The lower detector offers the type of contrast which we all see i=
n
our conventional single detector SEM, SE+BSE. The up side is that the BS=
E
contribution to this image results in far fewer charge problems.

3. Add a BSE detector to such a system, and you can move in any
direction "pure" SE or SE + BSE or pure BSE. Drop the kV and the BSE
becomes even more interesting as the volumes of material involved almost
mimic SE volumes.

All of you interested in image formation should be very jealous of someon=
e
who will be able to get his hands on an instrument like this; 100% scienc=
e!

When testing instruments of this type be very careful to select specimens=

that do look different at different kV. Gold on carbon is in my mind a
very poor performance standard for this type of instrument. It is often
unable to pull out the performance differences between 5 and 15kV. This
does not demonstrate a fantastic instrument but a poor test specimen.

You lucky man!

Steve Chapman

Senior Consultant E.M.
Protrain, 16 Hedgerley, Chinnor, Oxford OX9 4TN, England.
Tel & Fax 44 (0)1844 353161
Web Site - http://ourworld.compuserve.com/homepages/protrain
For Consultancy and Courses in Electron Microscopy World Wide





From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Sat, 5 Dec 1998 07:03:42 -0500
Subject: Low voltage Backscattered Imaging in FEG-SEMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Scott

The people I would contact in the FEG + BSE area are Geoff Richards and
Iolo ap Gwynn. They are very active in the biological side of BSE imagin=
g.


=46rom by own experiences you have no substitute for beam current, you ne=
ed
the ability to drive the gun at least up to 20uA, you will find some
manufactures even have 40uA if you know where to get at it!

Try Iolo ap Gwynn on iag-at-aber.ac.uk

Steve Chapman
Senior Consultant E.M.
Protrain, 16 Hedgerley, Chinnor, Oxford OX9 4TN, England.
Tel & Fax 44 (0)1844 353161
Web Site - http://ourworld.compuserve.com/homepages/protrain
For Consultancy and Courses in Electron Microscopy World Wide





From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Sat, 5 Dec 1998 11:34:27 -0500
Subject: That Game!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi to all my friends in South Africa,

Well after a great Conference and some foul weather what better than a go=
od
game of rugby?

We have to say that a really GREAT team were beaten on the day by a very
good team. Sorry you did not break the record (although your TV
commentator would have been moaning about the ref as he always seems to d=
o)
but you ALL have to admit that we beat you fair and square?

Keep smiling?

Shoestring





From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sat, 05 Dec 98 11:49:06 -0500
Subject: TEM Grids: "Dull vs Shinny"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Dr. Elaine Humphrey wrote:
============================================
We have recently had a number of users who have come from different places
in the world, and use different conventions for whether sections/formar go
on the shiny side of the grid or the dull side.

This lab uses the dull side, and has considered that the convention. If it
ain't broke don't fix it.

As long as the user is consistent with his/her own grids, there is no
problem, but when looking at collegues grids from another lab, it can lead
to a few annoyances.

We have been making our own formvar or pioloform-coated grids and putting
the formvar or pioloform on the dull side. One of my users has bought made-
up-grids where the coating was on the shiny side. At least two other labs
here use the shiny side for sections.

Is there a difference to the sides of the grid that makes one side better
than the other?
=====================================================
This is certainly the most often asked-for question about TEM grids, at
least for our firm, from customers writing or calling in for "advice" .

After being involved for more than thirty years with this question, I have
come to the following conclusions. I present them here for whatever peer
review might be forthcoming:

a) It seems to be about 50/50 (non-scientifically determined) on a
worldwide basis as to whether users prefer dull or shinny.

b) While the "dull" side has sharper "protrusions" and in theory at least
should initiate tears and cracks more readily than the shinny side, I am not
sure we ourselves have ever found evidence that is the case. Those
expressing preferences for the dull side tend to claim better adhesion
because of the larger surface area available for adhesion. So we have
ourselves given up on using our own good common sense to rationalize which
side is "better". We think that one can, in general, get comparable results
from either side.

c) However, not all grids are manufactured using the same processes and the
conclusions drawn from grids from one source might not necessarily be
applied to grids from other sources. For example, anyone who has used
different "brands" of grids knows that there is a difference between dull
and shinny side discrimination. We have even found customers who have
expressed a preference for grids showing the least dull/shinny difference so
it is not necessarily even a case of good vs. bad.

d) We ourselves believe that the real point of the dull vs. shinny side is
to quickly and easily discriminate between which side is "up" and which side
if "down" since many samples are not easily seen with the naked eye. Some
percentage of the population develop a level of color-blindness, and this is
also true as we get older, and as that happens, there also seems to be a
loss in ability to discriminate between dull vs. shinny. Hence, some grids
are actually purchased on that basis alone, without regard to any other
considerations. From that perspective it does not matter which side is
being used, just so one is consistent and always using the same side for
their specimens.

e) Paul Webster raises an interesting question relative to there being an
asymmetry (he called it a polarity) to the slot grids. We ourselves have
never looked at that aspect of the issue before. But if it is correct, then
it would probably also apply to the use of aperture grids as well, since
they are made under identical conditions. In general, it is the shinny side
that is the one in contact with the "glass master" during the grid making,
and the dull side is the one that "grows" out into the plating bath. In any
case the point is well made and this will encourage us to see what could be
done to reduce the asymmetry present in slot grids.

f) All of my comments and I presume those of the others are directed
specifically for those grids that are electro deposited, that is, Cu, Ni,
and Au and such comments would not be applicable to grids made by other
methods (for example, chemical etching, to produce grids from Be or W, etc.)




From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Sat, 05 Dec 98 11:49:12 -0500
Subject: New listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Dr. Rik Brydson wrote:
=====================================
Some information on a new listserver recently set up. I encourage you to
join Please display in your institution.

Many thanks,
================================================
It is not clear to me how this new listserver would in some way contribute
to the objectives as outlined, in ways not possible via the existing
listserver sponsored by the Microscopy Society of America. Right now, if
someone has a question, they need post it in only one place, if the "places"
were fragmented, to get the same full coverage, one would have to place it
in more than one place.

And for those who felt they wanted to subscribe to more than one listserver,
they are going to be reading the same postings more than one time.

I would cast my vote that the system we have "ain't broken" and therefore we
shouldn't try to fix it!

Chuck



===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: www.2spi.com
############################
==================================================





From: Ritchie Sims :      r.sims-at-auckland.ac.nz
Date: Sun, 6 Dec 1998 06:34:12 GMT+1200
Subject: Re: Query: How much better: WDS than EDS for SODIUM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



But the higher currents needed might upset your samples, mightn't
they?

rtch


} Jacob Bastacky wrote:
} }
} } We are considering looking for a Wavelength-Dispersive Spectrometer to
} } improve our detection of sodium in biological samples, reasoning that the
} } better energy resolution of the detector will result in narrower, taller
} } peaks with the same background level, yielding improved peak-to-background
} } ratios.
}
} This is correct.
} }
} } Our samples are frozen hydrated and dried biological tissues; we would like
} } to be able to differentiate 50mMolar from 150mMolar sodium (0.25 and .75%
} } by weight in hydrated samples).
} }
} This should be pretty straight-forward.
} }
} } Might we expect a significant improvement with WDS?
} }
} Yes.

Ritchie Sims Phone : 64 9 3737599 ext 7713
Department of Geology Fax : 64 9 3737435
The University of Auckland email : r.sims-at-auckland.ac.nz
Private Bag 92019
Auckland
New Zealand





From: crow-at-aloha.net (MHC)
Date: Sat, 5 Dec 1998 16:50:31 -1000 (HST)
Subject: looking for sliding microtome/rotary blades/instructions for blade sharpener

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are a non-profit botanical garden conducting plant anatomy research. We
can pay for the shipping of the following items:

We are looking for a Riechert Sliding Microtome & knives from
anyone/institution wishing to dispose of one.

We are also looking for non-disposible blades for a Spencer 820 rotary microtome

& just the INSTRUCTIONS for a Sensaur microtome knife sharpener:
Aloe - Cat # V60041; Model # SH 67-R; Serial # 1107; 115 Vac 50/60 C 5A

Please reply to my email: crow-at-aloha.net
Thanks very much in advance

_______________________________________________________________
mattie
email: crow-at-aloha.net
________________________________________________________________







From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Sun, 6 Dec 1998 05:49:37 -0500
Subject: Nester - Sorry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry mixed up my MSA and MSSA servers.

Steve





From: Nestor J. Zaluzec :      zaluzec-at-Sparc5.Microscopy.Com
Date: Sun, 6 Dec 1998 13:42:25 -0500
Subject: Administrivia: Archives Updated & M&M 99 Meeting

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues,

Sorry, I just realized how far behind I was in updating the Microscopy
Listserver Archives.

The archives are now current through November 31, 1998,
and can be found at the MSA WWW Site

http://www.msa.microscopy.com/MicroscopyListserver/MicroscopyArchives.html

In addition, information about Microscopy & Microanalysis 1999 Meeting
is now on-line at

http://www.microscopy.com/MSAMeetings/MMMeeting.html

Your Friendly Neighborhood SysOp

Nestor







From: Melvyn Dickson :      M.Dickson-at-unsw.edu.au
Date: Mon, 07 Dec 1998 07:48:01 +1100
Subject: Re: grids dull or shiny?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The bars on the shiny side have sharper edges and smoother surfaces than
the bars on the dull side.

Check this out with your SEM

The argument for using the dull side for the films is that there is less
risk of tearing at the edge of the bar and perhaps better adhesion.




*****************************************************
Mel Dickson,
Director.
Electron Microscope Unit,
University of New South Wales.
Sydney NSW 2052 Australia

Phone (+612) 9385-6383
Fax (+612) 9385-6400
Website {http://srv.emunit.unsw.edu.au}
*****************************************************





From: Paul Webster :      pwebster-at-mailhouse.hei.org
Date: 06 Dec 98 14:22:01 -0800
Subject: Re>TEM Grids - dull v shiny

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


One final thought on this:

Interference colors, which reveal the location of sections on the grid, =
are more easily seen on a dull background. Interference colors also play =
an important role in the evaluation of support film thickness during the =
final, drying step in cryosectioning. There is a reason for choosing one =
side after all.

Regards, =

Paul Webster, Ph.D
House Ear Institute
2100 West Third Street
Los Angeles, CA 90057
phone:213 273 8026
fax: 213 413 6739
e-mail: pwebster-at-hei.org
http://www.hei.org/htm/apw.htm






From: MicroToday-at-aol.com
Date: Sun, 6 Dec 1998 17:56:54 EST
Subject: Microscopist Salary Survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by imo22.mx.aol.com (IMOv18.1) id 8BYIa17651
for {microscopy-at-sparc5.microscopy.com} ; Sun, 6 Dec 1998 17:56:54 -0500 (EST)
Message-ID: {d2de1999.366b0bb6-at-aol.com}


Readers,
Due to the considerable number of request, we have decided to publish the
results of our recent microscopist salary survey. However, do to the
tabulation, I regret that I do not know how to publish it on this medium.
The full survey results will be in the December issue of Microscopy Today -
which will hit the P.O. on 17 December. If you have provided your salary data
and wish an earlier copy, send me your fax number and I will be pleased to fax
you a copy.
Don Grimes, Microscopy Today
P.S. #1: A surprise (to me) in the survey! The average yearly salary of all
457 microscopists reported was $51,676. The average yearly salary of all 268
male microscopists was $55,641 - or up some 7.7 % from the total average.
But, the average yearly salary of all 189 female microscopists was $46,053 -
or down some 10.9 % from the total average. No conclusion suggested!
Interesting?
P.S. #2: My sincere wishes for a nifty set of holidays to all - particularly,
Nestor, to you!





From: Jim J Darley :      jim-at-proscitech.com.au
Date: Mon, 7 Dec 1998 11:44:49 +1100
Subject: RE: grids dull or shiny?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Elaine:
Years ago I read a brief SEM study which looked at the
attachment of films and sections and concluded that the
dull side was slightly better. However, the grid bars edges
are a little more rounded on the dull side and supporting
films or sections would tend to go a bit more up and down
with that rounding.
I always used the dull side, but what matters is uniformity
in one working group.
Cheers
Jim Darley

ProSciTech Microscopy
PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 7 4774 0370 Fax: +61 7 4789 2313
Great microscopy catalogue, 500 Links, MSDS, User Notes
********************** www.proscitech.com.au *****



On Saturday, December 05, 1998 9:32 AM, Elaine Humphrey
[SMTP:ech-at-unixg.ubc.ca] wrote:
}
}
----------------------------------------------------------
} -------------.
}
}
} Hello all
} We have recently had a number of users who have come from
} different places
} in the world, and use different conventions for whether
} sections/formar go
} on the shiny side of the grid or the dull side.
}
} This lab uses the dull side, and has considered that the
} convention. If it
} ain't broke don't fix it.
}
} As long as the user is consistent with his/her own grids,
} there is no
} problem, but when looking at collegues grids from another
} lab, it can lead
} to a few annoyances.
}
} We have been making our own formvar or pioloform-coated
} grids and putting
} the formvar or pioloform on the dull side. One of my
users
} has bought
} made-up-grids where the coating was on the shiny side.
At
} least two other
} labs here use the shiny side for sections.
}
} Is there a difference to the sides of the grid that makes
} one side better
} than the other?
} Elaine
}
}
} Dr. Elaine Humphrey
} Biosciences Electron Microscopy Facility
} University of British Columbia
} 6270 University Blvd, mail-stop Botany
} Vancouver, BC
} CANADA, V6T 1Z4
} Phone: 604-822-3354
} FAX: 604-822-6089
} e-mail: ech-at-unixg.ubc.ca
}
}






From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Mon, 7 Dec 1998 11:28:52 +0000 (BST)
Subject: Re: Query: How much better: WDS than EDS for SODIUM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jacob:

Yes, WDS will be better for sodium but the trade off you have to make is
that you will probably have to put in more beam current to get
sufficient signal. A windowless or an atmosphere thin window (100nm)
should work okay. Depends what else is in the spectrum. I would try it
on a known standard. I think you are dead in the water with a beryllium
window, too many of the sodium x-ray photon would get absorbed.

Best Wishes for the Christmas Season

Look forward to seeing you next summer. Shirley and I will be on the
West Coast.

Patrick

On Thu, 3
Dec 1998, Jacob Bastacky wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} We are considering looking for a Wavelength-Dispersive Spectrometer to
} improve our detection of sodium in biological samples, reasoning that the
} better energy resolution of the detector will result in narrower, taller
} peaks with the same background level, yielding improved peak-to-background
} ratios.
}
} Our samples are frozen hydrated and dried biological tissues; we would like
} to be able to differentiate 50mMolar from 150mMolar sodium (0.25 and .75%
} by weight in hydrated samples).
}
} Our present EDS detector has a semi-thin window and is doing better than
} average for sodium and we would prefer not going to an ultrathin window or
} windowless detector.
}
} Might we expect a significant improvement with WDS?
}
} Does anyone know of a spectrometer that might be donated?
}
} Your comments and suggestions will be appreciated.
}
} Thanks,
}
} Jacob
}
} Jacob Bastacky, M.D.
} Room 116 Donner Laboratory
} Lawrence Berkeley Laboratory
} University of California
} Berkeley, California 94720
} Telephones: 510.486.4606 office, 510.486.4605 lab, 510.845.8031 fax
} email: sjbastacky-at-lbl.gov
}
}
}
}






From: Rosemary A. Walsh :      rw9-at-psu.edu
Date: Mon, 7 Dec 1998 07:38:51 -0500
Subject: EDS to monitor ion uptake in polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,
I am seeking your guidance on
using EDS to quantitate the ion uptake in
homogeneous polymers (cast at a uniform
thickness). Any references would be
appreciated as well.
Rosemary







From: Barbara Weyn :      ingridb-at-uia.ua.ac.be
Date: Mon, 7 Dec 1998 07:41:39 -0500
Subject: Lipstick

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have to make 10 u sections of lipstick.
Can somebody help us?







From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 12/4/98 2:44 PM
Subject: Re: Low voltage Backscattered Imaging in FEG-SEMs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


I would like to hear from microscopists who have used FEG-SEM's and
particularily the semi-in-lens models for biological applications. I would
also appreciate info on recent papers where this instrumentation was
critical to the biological research reported. Also any recommendations as to
ideal samples for testing these microscopes for biological application?
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057
--------------------------------------


I am evaluating FEG semi-in-lens SEMs for possible purchase. I am
interested in BSE imaging at low acc. voltages ( {3kV). I would like to
get
some feedback as to people's experiences with using BSE mode at low
voltages. Specifically, what kind of detectors are being used and what
kind
of results? What is the lowest acc. voltage that you can reasonably use
and
still get good results.
Thanks,



RFC822 header
-----------------------------------

RECEIVED: from SF_Database by POP_Mailbox_-1299317290 ; 04 DEC 98
21:00:06 UT
Received: from SPARC5.MICROSCOPY.COM by mailcenter.btny.purdue.edu
with SMTP (QuickMail Pro Server for MacOS 1.0.3); 04 DEC 98
20:59:50 UT
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com
(8.6.11/8.6.11) id NAA27632 for dist-Microscopy; Fri, 4 Dec 1998 13:50:05 -0600
Received: from no_more_spam.com (Sparc5 [206.69.208.10]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id NAA27623 for
"MicroscopyFilteredEmail-at-msa.microscopy.com"; Fri, 4 Dec 1998 13:49:34 -0600
Received: from gateway.ppg.com (gateway.ppg.com [199.221.65.2]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id NAA27616 for
{microscopy-at-sparc5.microscopy.com} ; Fri, 4 Dec 1998 13:49:22 -0600
Received: by gateway.ppg.com id PAA26769
(SMTP Gateway for microscopy-at-sparc5.microscopy.com);
Fri, 4 Dec 1998 15:01:40 -0500
Message-Id: {199812042001.PAA26769-at-gateway.ppg.com}
Received: by gateway.ppg.com (Protected-side Proxy Mail Agent-1);
Fri, 4 Dec 1998 15:01:40 -0500
From: "Walck. Scott D." {walck-at-ppg.com}
To: Micro {microscopy-at-Sparc5.Microscopy.Com}
Subject: Low voltage Backscattered Imaging in FEG-SEMs
Date: Fri, 4 Dec 1998 14:44:00 -0500
Mime-Version: 1.0
X-Mailer: Internet Mail Service (5.5.2232.9)
Content-Type: text/plain
Errors-to: Microscopy-request-at-sparc5.microscopy.com








From: Scott Holt :      holt_scott-at-CompuServe.COM
Date: Mon, 7 Dec 1998 09:37:42 -0500
Subject: Al-Si alloy pullout

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by arl-img-4.compuserve.com (8.8.6/8.8.6/2.16) id JAA07014
for microscopy-at-sparc5.microscopy.com; Mon, 7 Dec 1998 09:38:06 -0500 (EST)


Dear Dorrance,

To start with, I should mention that I am an interested party since =

BUEHLER is a major supplier of polishing equipment, consumables,
and technical support for the grinding/polishing industry and has been
for over 60 years.

To my point: Although I agree with Gary Leichty's point about using
diamond =

vs. silicon carbide papers, it is for different reasons. Silion Carbide =
is
a brittle
abrasive with a high coefficient of friction compared to diamond. Due to=

this fact,
and as a result of being fixed to a paper backing material, SiC tends to
'plow' material
during grinding operation. It also tends to fracture, creating smaller
particles that
can embed in ductile phases. Since the SiC abrasive is rigidly fixed to
the paper backing
material, when it 'plows' into the metallic phase of your alloy, it
eventually contacts
a hard precipitate and plucks it out.

Diamond films also 'plow' ductile phases because the abrasive is fixed to=
a

backing material. In this case, as opposed to SiC papers, the resin bind=
er
of the film is
weaker and allows the diamond to pull free of the film and stay embedded =
in
the sample. =

Although proper lubrication will help, diamond films are not the answer i=
n
your case. =


A better suggestion might be to section a disk from the bulk using a thin=

diamond blade, making sure that the sample is near final thickness. Then=

use one or more sizes of diamond paste product on a porous pad for removi=
ng

damage created by the blade. The waxy carrier of paste products along wi=
th
embedding of the diamond in the pad will tend to eliminate embedding in
your sample.

Finally, using a low napped polishing cloth with a colloidal silica produ=
ct
(High
pH) will produce a smooth, low deformation surface in which no embedding =

exists, and in which both the aluminum and precipitate phases are
relatively
coplanar.

Please contact me directly if you have additional questions.

Best regards,

Scott D. Holt
BUEHLER, Ltd.
PO Box 1
41 Waukegan Rd.
Lake Bluff, IL 60044
(847)295-4546
****** http://www.buehlerltd.com ******

*************************************************************************=
**
********

Dorrance wrote:

} Dear All,
}
} I've been batting my head against my Fischione Jet Polisher trying to
figure
} this one out! Please help if you can. My original material is full of=

} lovely voids and pits, Si particles are around 10 microns, and at a
} thickness of 150 to 200 microns I have experienced some holes that go a=
ll
} the way through the material. If I try initial thinning of the sample =
on
} less than 500 grit SiC paper I get even more particle pullout...Any
ideas?
}
} Thanks in advance for any help.
}
} Dorrance





From: Marti, Jordi :      Jordi.Marti-at-alliedsignal.com
Date: Saturday, December 05, 1998 11:49AM
Subject: New listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I tend to agree with Chuck.

So far the listserver has been a good source of infomation. I think we
would all lose if we splinter into smaller listservers. If anything I
would consider using better (i.e. clearer) Subject Identifiers. For
example, one could use headers such as Bio (for Biological), Mat (for
Materials ) and Gen. (for general ) subjects. This would make it easier for
a reader to decide if he/she wants to open a given posting. I believe
something like this was proposed by others in the past.

We also get much useful information from vendors (I am not a vendor) and
if we were to split this would pose increased problems for both vendors
as well as us microscopists in terms of the time spent searching and
posting messages.

In short, I think we would lose more than what we would gain.

Jordi Marti


----------
} From: Garber, Charles A.
To: MICROSCOPY BB
-----------------------------------------------------------------------.


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Dr. Rik Brydson wrote:
=====================================
Some information on a new listserver recently set up. I encourage you to
join Please display in your institution.

Many thanks,
================================================
It is not clear to me how this new listserver would in some way contribute
to the objectives as outlined, in ways not possible via the existing
listserver sponsored by the Microscopy Society of America. Right now, if
someone has a question, they need post it in only one place, if the "places"
were fragmented, to get the same full coverage, one would have to place it
in more than one place.

And for those who felt they wanted to subscribe to more than one listserver,
they are going to be reading the same postings more than one time.

I would cast my vote that the system we have "ain't broken" and therefore we
shouldn't try to fix it!

Chuck



===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: www.2spi.com
############################
==================================================






From: savost-at-sens.ief.spb.su (Savostjanov G.A.)
Date: Mon, 7 Dec 98 17:51:22 +0300
Subject: LM,SEM - need consultation and cooperation in 3-D reconstr. of epithelial sheets histoarchitecture

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Colleagues
Being aware of your interest in 3-D reconstruction and
visualization of biological object structures, I wonder if you
deal with the problem of 3-D epithelial sheets topology
reconstruction? If it is so can you send me your papers on this
subject? And do you know anybody who work in this field?
My own interest is the 3-D histoarchitecture of epithelial
sheets in normal development and pathology.
In this connect I am addressing you with the following
message.
As you know, all current attempts to reconstruct a 3-D
structure of biological tissues using a serial sections
encounter with a typical difficulty. It is that tissue
deformation and cell proliferation make a cells shape variable,
diverse their adjacency and give a strongly "noised" pictures.
So the results obtained give no way to understanding tissue
topology and do not permit to deduce the law of 3-D tissue
organization.

To overcome the difficulties we have developed the new
approach to 3-D reconstruction of cell sheet structures. It is
based on the pioneering concept of tissues modular structure and
consists of derivation of topological and geometrical models of
epithelial spatial organization and their experimental
verification. Contrary to the usual tissue reconstruction on a
set of serial sections this approach allows to use their minimum
and to carry out a more exact restoration of 3-D epithelial
structure with less money and time. The approach has enabled us
to get the data priority on the laws of spatial organization of
simple, pseudostratified and stratified epithelia, as well as to
predict and find several earlier unknown topological variants of
their histoarchitectonics. The approach has also allowed to
describe such new properties of epithelial layers as translation
symmetry and stoichiometry.

The results makes a basis for structural histology as a
addition to modern structural biology. The approach allow to
investigate the more complex topology variants of
pseudostratified and stratified epithelia, to find a set of new
informative tissue characteristics suitable for diagnostic and
also to give the opportunity to predict the changes of tissues
in normal development and their transformation in pathology,
particularly in malignant growth.

If it is interesting for you, I am ready to consider the
possibility of our cooperation.

Yours sincerely,
Dr. G.A. Savostyanov.
--
| E-mail: savost-at-ief.spb.su | Gennady A. Savostyanov |
| Fax +7 (812) 5523012 |Sechenov Inst. of Evolutionary Physiology and |
| office +7 (812) 5523090 | Biochemictry Russian Academy of Science, | |
| home +7 (812) 5100052 |44 Thores av., 194223, St.Petersburg, Russia |





From: Barbara Foster :      mme-at-map.com
Date: Mon, 07 Dec 1998 11:51:58 -0500
Subject: Re: Lipstick

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ingrid,


What is the purpose of your having to section? If it is to determine
pigment distribution, have you tried confocal? As long as there is not
titantium dioxide as a filler (it tends to act as a mirror), confocal may
work, saving you lots of aggrevation.


If you have to section, talk to Wasi Ahmed at Buehler Instruments here in
the States. He had sectioned everything from light bulbs to the ash on
the end of a cigar! Try them at 847-295-6500.


Best regards,

Barbara Foster

Consortium President

{color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
..Educating microscopists for greater productivity.


{/color} 125 Paridon Street Suite 102 Springfield, MA 01118

PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

Visit our web site { {http://www.MME-Microscopy.com/education}

******************************************************

{bigger} {bigger} MME {/bigger} {/bigger} is America's first national
consortium dedicated to

customized on-site training in all areas of

microscopy, sample preparation, and image analysis {bigger} .

{/bigger}





At 07:41 AM 12/7/98 -0500, Barbara Weyn wrote:

} ------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

} -----------------------------------------------------------------------.

}

}

} We have to make 10 u sections of lipstick.

} Can somebody help us?

}

}

}

}

}







From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Mon, 7 Dec 1998 09:04:37 -0800 (PST)
Subject: Re: Microscopist Salary Survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} P.S. #1: A surprise (to me) in the survey! The average yearly salary of all
} 457 microscopists reported was $51,676. The average yearly salary of all 268
} male microscopists was $55,641 - or up some 7.7 % from the total average.
} But, the average yearly salary of all 189 female microscopists was $46,053 -
} or down some 10.9 % from the total average. No conclusion suggested!
} Interesting?

Not interesting, just sad. No matter how good we are or how hard we try,
women still are considered second class citizens and our salaries reflect
that. We're used to, but not happy about, being paid less almost 100% of
the time.


Paula :-(

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu
phone: 510-642-2085
fax: 510-643-6207
http://biology.berkeley.edu/EML







From: Gerhard Frank :      frank-at-ww.uni-erlangen.de
Date: Mon, 07 Dec 1998 18:07:11 +0100
Subject: Magnification calib in STEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All.
How can I calibrate the maginfication in a scanning transmission
electron microscop for } 100.000x? Up to this mag one can use cross
grating replica, but beyond?
Thank you very much.
--
Gerhard Frank
UNIVERSITAET ERLANGEN-NUERNBERG Phone: +49 9131 85 28606
Institut fuer Werkstoffwissenschaften VII Fax: +49 9131 85 28602
Cauerstr. 6, D-91058 Erlangen, Germany frank-at-ww.uni.erlangen.de
*****************************
* NEW Phone and Fax numbers!*
*****************************





From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Mon, 7 Dec 1998 09:09:06 -0800 (PST)
Subject: Re: Microscopist Salary Survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



P.S. #1: A surprise (to me) in the survey! The average yearly salary of all
457 microscopists reported was $51,676. The average yearly salary of all 268
male microscopists was $55,641 - or up some 7.7 % from the total average.
But, the average yearly salary of all 189 female microscopists was $46,053 -
or down some 10.9 % from the total average. No conclusion suggested!
Interesting?


Not interesting, just sad. No matter how good we are at what we do or how
hard we try, women still are considered second class citizens and our
salaries reflect that. We're used to, but not happy about, being paid less
almost 100% of the time.


Paula :-(

p.s. everytime I try to reply to the
MicroToday-at-aol.com-at-sparc5.microscopy.com I get a bad address and the
message is thrown back, does anybody know why?

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu
phone: 510-642-2085
fax: 510-643-6207
http://biology.berkeley.edu/EML







From: William McManus :      billemac-at-bioserver.vsb.usu.edu
Date: Mon, 7 Dec 1998 10:17:02 -0700
Subject: RE: Lipstick

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




---------------------------------------.

Barabara Weyn wrote:
We have to make 10 u sections of lipstick.
Can somebody help us?

+++++++++
Try cryosectioning on a LM type cryostat.


William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920





From: Claypool :      pclypool-at-sgi.net
Date: Mon, 07 Dec 1998 12:57:01 -0500
Subject: Re:Listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



--------------D8DB0F645A0C698C249B773C
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

There's another option of course, there is a newsgroup titled
alt.sci.microscopy. My ISP seems to have a problem w/ this one though,
casue I lose all the threads in about a week.

It works basically the same was as this listserver, only you have the
option of reading or not reading messages, rather than having all
messages sent to you via email.

When i have a question or such, I'll usually post in both places.
For example, I posted my questions regarding the TN5500 here and there.
I recieved a few responses from there and many more from here. I am not
sure, however, how many people know of the newsgroup.

Well, thats my two pennies, hope it helps.

Ted Claypool
EPMA
RJ Lee Group

--------------D8DB0F645A0C698C249B773C
Content-Type: text/html; charset=us-ascii
Content-Transfer-Encoding: 7bit

{!doctype html public "-//w3c//dtd html 4.0 transitional//en"}
{html}
There's another option of course, there is a newsgroup titled alt.sci.microscopy. 
My ISP seems to have a problem w/ this one though, casue I lose all the
threads in about a week.
{p} It works basically the same was as this listserver, only you have the
option of reading or not reading messages, rather than having all messages
sent to you via email.
{p} When i have a question or such, I'll usually post in both places.
{br} For example, I posted my questions regarding the TN5500 here and there. 
I recieved a few responses from there and many more from here.  I
am not sure, however, how many people know of the newsgroup.
{p} Well, thats my two pennies, hope it helps.
{p} Ted Claypool
{br} EPMA
{br} {a href="http://www.rjlg.com"} RJ Lee Group {/a} {/html}

--------------D8DB0F645A0C698C249B773C--






From: shAf :      mshaf-at-darkwing.uoregon.edu
Date: Mon, 7 Dec 1998 10:27:33 -0800
Subject: RE: Microscopist Salary Survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Paula writes ...

} ...
}
}
}
} ...
}
}
} Not interesting, just sad. No matter how good we are at what
} we do or how
} hard we try, women still are considered second class citizens and our
} salaries reflect that. ...
} ...

I believe you are correct in many circumstances, but I think
you are wrong to make such a conclusion in this case when the
"average experience" wasn't reported with the "average salary" ...
also not reported with respect to gender; minimum and maximum
salaries and minimum and maximum degrees of experience ...
(leastwise in this e-mail thread ... I have yet to see the survey).
For example. I didn't take part in this survey for my own reasons
and I receive a lesser salary too with 20years experience.
Personally, I have a lesser salary because I chose this working
place and living environment, and didn't believe my salary
deserved to be rated with general electron microscopists. My lesser
salary would then come under a heading of "aggression for $$$", and
I do believe men are more aggressive when it comes to salary demands.
Regarding "2nd class citizens" I think things are more complex
than you know and not as bad as you imply, but I will agree ... for
the same experience salaries should be equal, and I don't think
were there yet either.

cheerios, shAf

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - ICQ 210524
Geological Science's Electron Probe Facility - University of Oregon
mshaf-at-darkwing.uoregon.edu - http://darkwing.uoregon.edu/~mshaf/







From: Bob Wise :      wise-at-vaxa.cis.uwosh.edu
Date: Mon, 07 Dec 1998 12:32:52 -0600 (CST)
Subject: Salary survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} P.S. #1: A surprise (to me) in the survey! The average yearly salary of all
} 457 microscopists reported was $51,676. The average yearly salary of all 268
} male microscopists was $55,641 - or up some 7.7 % from the total average.
} But, the average yearly salary of all 189 female microscopists was $46,053 -
} or down some 10.9 % from the total average. No conclusion suggested!
} Interesting?

Years-in-grade (or seniority) and rank have very large and statistically
significant impacts on salary. Were the gender data normalized to account
for this? If not, then no conclusion is possible.

Bob


Dr. Robert R. Wise
Department of Biology and Microbiology
University of Wisconsin-Oshkosh
Oshkosh, WI 54901

(920) 424-3404 tel
(920) 424-1101 fax
wise-at-uwosh.edu
www.uwosh.edu/departments/biology/wise/wise.html







From: Barbara Foster :      mme-at-map.com
Date: Mon, 07 Dec 1998 13:38:17 -0500
Subject: RE: New listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,


Usually I try to stay out of the "political" discussions, but I would
like to put my two cents in on this one.


My initial training in microscopy came from the Royal Microscopical
Society in England. One thing which differentiated their approach from
more typical US courses was their cross-disciplinary perspective:
electron microscopists talked to light microscopists and materials
scientists learned new staining approaches from biologists, etc. We all
benefit from the diverse postings on the MSA listserver. I agree with
Jordi, that headers would be helpful when our mailboxes get too full, but
I, for one, learn as much as possible about "what the other guy is doing"
so that I can add to my own experience base.


By the way: as an avid "industry watcher", I can tell you with certainty
that the current trend in professional societies is toward partnering
with sister-disciplines. The fact that our annual meeting is now M&M,
not just MSA, should be evidence enough. Let's keep the splintering to a
minimum and use headers, instead, to divvy up the info.


Hope this is helpful.

Barbara Foster

Consortium President

{color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
..Educating microscopists for greater productivity.


{/color} 125 Paridon Street Suite 102 Springfield, MA 01118

PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

Visit our web site { {http://www.MME-Microscopy.com/education}

******************************************************

{bigger} {bigger} MME {/bigger} {/bigger} is America's first national
consortium dedicated to

customized on-site training in all areas of

microscopy, sample preparation, and image analysis {bigger} .

{/bigger}






At 08:10 AM 12/7/98 -0700, Marti, Jordi wrote:

} ------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

} -----------------------------------------------------------------------.

}

}

} I tend to agree with Chuck.

}

} So far the listserver has been a good source of infomation. I think
we

} would all lose if we splinter into smaller listservers. If anything
I

} would consider using better (i.e. clearer) Subject Identifiers. For

} example, one could use headers such as Bio (for Biological), Mat (for

} Materials ) and Gen. (for general ) subjects. This would make it easier
for

} a reader to decide if he/she wants to open a given posting. I
believe

} something like this was proposed by others in the past.

}

} We also get much useful information from vendors (I am not a vendor)
and

} if we were to split this would pose increased problems for both
vendors

} as well as us microscopists in terms of the time spent searching
and

} posting messages.

}

} In short, I think we would lose more than what we would gain.

}

} Jordi Marti

}

}

} ----------

} } From: Garber, Charles A.

} To: MICROSCOPY BB

} Subject: New listserver

} Date: Saturday, December 05, 1998 11:49AM

}

}
------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America


} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

}
-----------------------------------------------------------------------.

}

}

} -- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

}

} Dr. Rik Brydson wrote:

} =====================================

} Some information on a new listserver recently set up. I encourage you
to

} join Please display in your institution.

}

} Many thanks,

} ================================================

} It is not clear to me how this new listserver would in some way
contribute

} to the objectives as outlined, in ways not possible via the existing

} listserver sponsored by the Microscopy Society of America. Right now,
if

} someone has a question, they need post it in only one place, if the
"places"

} were fragmented, to get the same full coverage, one would have to place
it

} in more than one place.

}

} And for those who felt they wanted to subscribe to more than one
listserver,

} they are going to be reading the same postings more than one time.

}

} I would cast my vote that the system we have "ain't broken" and
therefore we

} shouldn't try to fix it!

}

} Chuck

}

}

}

} ===================================================

} Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400

} President 1-(800)-2424-SPI

} SPI SUPPLIES FAX: 1-(610)-436-5755

} PO BOX 656 e-mail: cgarber-at-2spi.com

} West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com

}

}

} Look for us!

} ############################

} WWW: www.2spi.com

} ############################

} ==================================================

}

}

}

}







From: Jonathan Barnard :      J.Barnard-at-bristol.ac.uk
Date: Mon, 07 Dec 1998 19:06:49 +0000
Subject: Re: (STEM) calibration.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


What about STEM lattice imaging?

I would suggest using a thin 6H SiC specimen in (0001) orietation. 6H -
SiC has a lattice spacing of c=1.511 nm , so try to image these lattice
fringes in the STEM, but you may find that the collection angle of your
objective aperture to be too big i.e. you may have three or four beams,
but take several images at slightly different defocus to see what the
minimum spacing is between them.
Kinematically the (000m) - all m other than m=6n (integer n) beams are
forbidden, but they are strong (dynamically) close to and on certain
zone axes along the (000n) Kikuchi bands.
Good Luck.
JOn
--
*****************************************
Jonathan Barnard

Microstructural Physics,
H.H.Wills Physics Laboratory,
University of Bristol,
Tyndall Avenue,
Bristol BS8 1TL.

0117 928 9000 ext 8750

*****************************************







From: Sjoplinh-at-aol.com
Date: Mon, 7 Dec 1998 14:09:50 EST
Subject: LM Glycerol Borate mount medium?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



The little info I have suggests that Glycerol Borate might be a nearly ideal
mounting medium for my purposes. But more info seems hard to find.

It was once available commercially (from ?Glyco?) as "Aqua Resin". Alas, no
more. It has been recommended by Dr. McCrone himself. I thought it best to
try the list before asking Dr. M. or anyone at MRI.

Can this stuff be made by merely dissolving Borax in glycerin (+ water
initially?), or is it more complicated? Does anyone have any recipes? Any
tidbits of info will be appreciated.

Scott Harden





From: Darus, Mark :      DarusM-at-aerospace.bfg.com
Date: Mon, 7 Dec 1998 13:50:55 -0500
Subject: Subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Subscribe, thanks

DarusM-at-aerospace.bfg.com





From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Mon, 07 Dec 1998 13:56:22 -0600
Subject: Error in reply address

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


If you look closely at the address you will see a couple of "-at-" signs. If
you pull off the trailing "-at-sparc5.microscopy.com" then you should have a
valid address.

I have noticed the same problem on occasion and have so learned to
double-check the address before clicking the SEND button. If I forget and
get an error, then I know where to look for trouble.

I think Nestor may have explained why this happens. I forget the
explanation, and it is not always a problem. So I basically just try to be
careful.

Warren


At 09:09 AM 12/7/98 -0800, you wrote:
} p.s. everytime I try to reply to the
} MicroToday-at-aol.com-at-sparc5.microscopy.com I get a bad address and the
} message is thrown back, does anybody know why?






From: Douglas W. Darnowski :      darnowsk-at-staff.uiuc.edu
Date: Mon, 7 Dec 1998 14:15:49 -0500
Subject: Salary survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Paula, and others who have read her message as regards the salary survey,

The reply to the listserv:

"Not interesting, just sad. No matter how good we are or how hard we try,
women still are considered second class citizens and our salaries reflect
that. We're used to, but not happy about, being paid less almost 100% of
the time.


Paula :-(

Paula Sicurello"


is an example of sloppy thinking. One piece of summary statistical data
does not demonstrate anything about the treatment of a group as there may
be systematic factors which explain data such as was cited. After all,
there is more data, according to the authors of the survey, not yet put on
the listserv due to formatting questions.

After all, those who conducted the survey wrote previously that there were
not enough data points to justify the publication of the survey, in their
original opinion. The use of small samples to measure a group leads to
errors. The fact that those who conducted the survey brought up such a
point after having originally said that there was not enough data to
publish suggests that they may have been fishing for statements such as
Paula's. In my opinion, that is an irresponsible use of survey data.

A major problem today in the media, one which frustrates scientists such as
those who pay attention to this list, is the poor understanding of
statistics and their use in science, social or natural, held by reporters
and the general public. This can lead to poor decision making which can in
turn hurt scientific research in terms of both funding and public opinion.
As a good example, consider the nonsense about power lines and cancer.

The comment in question here is another example of such weak reasoning. If
scientists choose to think these questions through in such a manner, how
can we expect the general public to do otherwise?

Doug Darnowski
******************************************************************************
Douglas Darnowski
Department of Crop Sciences
384 ERML
1201 West Gregory Drive
University of Illinois
Urbana IL 61801
work ph: (217) 244-6150
fx: (217) 333-4777
home ph: (217) 356-6606
fx: (217) 356-4454
email: darnowsk-at-staff.uiuc.edu





From: Sara Miller :      saram-at-acpub.duke.edu
Date: Mon, 7 Dec 1998 15:04:51 -0500 (EST)
Subject: Re: TEM Grids: "Dull vs Shinny"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We routinely use the dull side, not only because of the increased
"stickiness" of the membrane to the rough surface, but also because if
you put the film on the shiny side and then coat it with carbon as we
always do, you have 2 dull sides. It is, thus, more difficult to
determine where the sample is when one flips over--particularly if you're
doing negative stains. This may not be the case for cryosections
in methyl cellulose or epoxy sections, but we always use the dull side
for these too to be consistent.

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735






From: Barbara Foster :      mme-at-map.com
Date: Mon, 07 Dec 1998 15:24:11 -0500
Subject: Re: LM,SEM - need consultation and cooperation in 3-D

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dr. Savostyanov,

Confocal microscopy offers a good solution to this problem. Also, you
might want to investigate the technology offered by Edge Scientific, both
their R400 (a full, real time 3D imaging system which can use Phase and
DIC) as well as their new Perceptra illuminating system. I will also
forward some other information to you under separate cover. Edge can be
reached by email at edgesci-at-primenet. In terms of confocal, there are a
number of choices, ranging from BioRad to Carl Zeiss, Leica, Olympus, and
Nikon. EG&G Wallac seems to be the newest player in the confocal field.
For these companies, I suggest you visit either the MSA web site at
www.msa.microscopy.com, MicroWorld at www.mwrn.com, or the Microscopy
Vendors' Database at www.kaker.com.

Hope this is helpful.

Best regards,








At 05:51 PM 12/7/98 +0300, Savostjanov G.A. wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
| |
} | home +7 (812) 5100052 |44 Thores av., 194223, St.Petersburg, Russia |
}
}
}






From: Goodhouse, Joseph :      jgoodhouse-at-molbio.princeton.edu
Date: Mon, 7 Dec 1998 15:51:19 -0500
Subject: Bio-rad MRC600 & Y2K

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Confocal users,
Below is an email I sent to Bio-rad regarding MRC600's and
Y2K. I have received no response from them regarding these issues. Does
anyone out there know the answers to these issues?
Dear Bio-rad,
I am the Confocal / Electron Microscopy Core Facility Manager
for the Department of Molecular Biology at Princeton University. I am
inquiring as to the Y2K status of our MRC600 Confocal Microscope System. We
have had this instrument since 1991, and it continues to be a valuable
instrument for this laboratory, and my
department. We have collected and archived over 150 gigabytes of image
data on optical disks with this instrument. This data needs to be
retrievable currently, as well as in the future. The instrument is
under service contract We are currently running the Comos7.0 program under
Windows 95 software. We are
also running macro programs out of SOM under COMOS version 6.03. These
macro programs are an essential tool for several of my investigators
imaging GFP (green fluorescent protein) in living organisms. Without
these macro programs to run the instrument, we would be unable to obtain
the information required for these experiments. We also run CAS 2.5 and
CAS 3.10 for converting images to tiff formats and for creating 24 bit
merged color tiff files for export to other programs.
I need to know
a) if this instrument will continue operating on and after
01-01-2000,
b) whether or not previously collected images will be read by
the program,
c) if we will continue to be able to run macros under the SOM
command shell,
d) what software and / or hardware upgrades will be required,
and when you will have them available?

Sincerely,

Info & Images at http://www.molbio.princeton.edu/confocal/

Joe Goodhouse
Confocal / EM Core Lab Manager
Department of Molecular Biology
Princeton University
jgoodhose-at-molbio.princeton.edu
609-258-5432







From: WPTT39A-at-PRODIGY.COM (DR ROBERT DAVIS)
Date: Mon, 7 Dec 1998 16:44:25, -0500
Subject: Purchase of a Microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have been looking for a binocular phase-contrast microscope during
the last year or so. I recently received an A. Daigger & Company,
Inc., Lincolnshire, Illinois catalogue which lists a reasonable
priced (1095.00) microscope. The microscope is a LABOROCON (TM) with
a 5 year warranty. The price seems reasonable, but I have no
experience with this product and don't know anyone who has. I will
appreciated hearing from anyone who has any experience with this
microscope. I would be interested in learning if anyone knows of
somone who handles second hand microscopes that I might contact. As
I am not a subscriber, please contact me directly at WPTT39A-at-prodigy.
com
Regards, Bob Davis





From: MelanieOwl-at-aol.com
Date: Mon, 7 Dec 1998 16:50:31 EST
Subject: Re: Salary survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Whether this particular study is statistically sound in terms of men's vs.
women's salaries, it is well documented (C&E News and others) that this is
still a problem in the scientific arena, even though things have improved from
years ago.
I happen to think it is quite humorous that whenever this issue comes up, in
any conversation or whatever, some man or men immediately jump up and start
defending themselves by claiming that this is not true, or even worse, ranting
about reverse discrimination that they claim to know all about and have
experienced! So defensive! I wonder why?

M. Behrens





From: Dr. Rachel Teitelbaum :      teitelba-at-aecom.yu.edu
Date: Mon, 7 Dec 1998 17:26:12 -0500 (EST)
Subject: the old boys network

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


So how does one explain the dominance of men in the field, as evidenced by
fewer tenured professors, etc., and I don't know if NIH breaks down its
grants this way, but there are certainly many fewer Howard Hughes
recipients who are female, and nobel laureates for that matter. I think
it is a rational deduction that there is much less respect for women in
science, and that despite the changing times, there is still an "old boys
network" in place. Surveys in the business field have certainly
demonstrated that women are often paid less despite similar
qualifications. I'm sorry that more people did not respond to the survey
to achieve statistical significance, as to how the results might have
changed with more responses.....

-Rachel






From: Gillian Bond :      gbond-at-nmt.edu
Date: Mon, 7 Dec 1998 15:36:35 -0700 (MST)
Subject: RE: New listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I would like to add my whole-hearted agreement. One of the things I value
about the MSA listserver is the ability to read postings from, and make
contacts with microscopists from a cross-disciplinary range of
backgrounds, particularly since one of my own research areas is
interdisciplinary in nature.

Gill Bond
Dept Materials & Met. Eng.
New Mexico Tech

On Mon, 7 Dec 1998, Barbara Foster wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear all,
}
}
} Usually I try to stay out of the "political" discussions, but I would
} like to put my two cents in on this one.
}
}
} My initial training in microscopy came from the Royal Microscopical
} Society in England. One thing which differentiated their approach from
} more typical US courses was their cross-disciplinary perspective:
} electron microscopists talked to light microscopists and materials
} scientists learned new staining approaches from biologists, etc. We all
} benefit from the diverse postings on the MSA listserver. I agree with
} Jordi, that headers would be helpful when our mailboxes get too full, but
} I, for one, learn as much as possible about "what the other guy is doing"
} so that I can add to my own experience base.
}
}
} By the way: as an avid "industry watcher", I can tell you with certainty
} that the current trend in professional societies is toward partnering
} with sister-disciplines. The fact that our annual meeting is now M&M,
} not just MSA, should be evidence enough. Let's keep the splintering to a
} minimum and use headers, instead, to divvy up the info.
}
}
} Hope this is helpful.
}
} Barbara Foster
}
} Consortium President
}
} {color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
} ..Educating microscopists for greater productivity.
}
}
} {/color} 125 Paridon Street Suite 102 Springfield, MA 01118
}
} PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
}
} Visit our web site { {http://www.MME-Microscopy.com/education}
}
} ******************************************************
}
} {bigger} {bigger} MME {/bigger} {/bigger} is America's first national
} consortium dedicated to
}
} customized on-site training in all areas of
}
} microscopy, sample preparation, and image analysis {bigger} .
}
} {/bigger}
}
}
}
}
}
}
} At 08:10 AM 12/7/98 -0700, Marti, Jordi wrote:
}
} } ------------------------------------------------------------------------
}
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
}
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
}
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
} } -----------------------------------------------------------------------.
}
} }
}
} }
}
} } I tend to agree with Chuck.
}
} }
}
} } So far the listserver has been a good source of infomation. I think
} we
}
} } would all lose if we splinter into smaller listservers. If anything
} I
}
} } would consider using better (i.e. clearer) Subject Identifiers. For
}
} } example, one could use headers such as Bio (for Biological), Mat (for
}
} } Materials ) and Gen. (for general ) subjects. This would make it easier
} for
}
} } a reader to decide if he/she wants to open a given posting. I
} believe
}
} } something like this was proposed by others in the past.
}
} }
}
} } We also get much useful information from vendors (I am not a vendor)
} and
}
} } if we were to split this would pose increased problems for both
} vendors
}
} } as well as us microscopists in terms of the time spent searching
} and
}
} } posting messages.
}
} }
}
} } In short, I think we would lose more than what we would gain.
}
} }
}
} } Jordi Marti
}
} }
}
} }
}
} } ----------
}
} } } From: Garber, Charles A.
}
} } To: MICROSCOPY BB
}
} } Subject: New listserver
}
} } Date: Saturday, December 05, 1998 11:49AM
}
} }
}
} }
} ------------------------------------------------------------------------
}
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
}
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
}
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
} }
} -----------------------------------------------------------------------.
}
} }
}
} }
}
} } -- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --
}
} }
}
} } Dr. Rik Brydson wrote:
}
} } =====================================
}
} } Some information on a new listserver recently set up. I encourage you
} to
}
} } join Please display in your institution.
}
} }
}
} } Many thanks,
}
} } ================================================
}
} } It is not clear to me how this new listserver would in some way
} contribute
}
} } to the objectives as outlined, in ways not possible via the existing
}
} } listserver sponsored by the Microscopy Society of America. Right now,
} if
}
} } someone has a question, they need post it in only one place, if the
} "places"
}
} } were fragmented, to get the same full coverage, one would have to place
} it
}
} } in more than one place.
}
} }
}
} } And for those who felt they wanted to subscribe to more than one
} listserver,
}
} } they are going to be reading the same postings more than one time.
}
} }
}
} } I would cast my vote that the system we have "ain't broken" and
} therefore we
}
} } shouldn't try to fix it!
}
} }
}
} } Chuck
}
} }
}
} }
}
} }
}
} } ===================================================
}
} } Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
}
} } President 1-(800)-2424-SPI
}
} } SPI SUPPLIES FAX: 1-(610)-436-5755
}
} } PO BOX 656 e-mail: cgarber-at-2spi.com
}
} } West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com
}
} }
}
} }
}
} } Look for us!
}
} } ############################
}
} } WWW: www.2spi.com
}
} } ############################
}
} } ==================================================
}
} }
}
} }
}
} }
}
} }
}
}
}






From: Sally Stowe :      STOWE-at-rsbs.anu.edu.au
Date: Tue, 8 Dec 1998 09:37:12 +1000 GMT
Subject: new listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My two cents -
Some of these issues came up when we started an Australasian
microscopy listserver a few years ago. I think there has to be some
specialisation, otherwise the global system will get clogged with
groups in Rekjavik, East Wapping and Te Awamutu announcing local
meetings etc.
We find a three-level system seems to work pretty well
- so we run a server for Canberra, and one for Australasia, and
its a commonsense decision whether to send a message to one of these
or to MSA.
So the MSA server doubles as the global as well as
the national US system, because it was first up, filled a need,
can handle the traffic, and because Nestor is willing to do the
work. For which much thanks, Nestor!
The establishment of more national servers outside the US is a
natural evolution, if they are dealing with purely regional issues.
But trying to split the "world-wide" conversation place into regional
groups would be pointless. Probably any schism is liable to be
self-healing anyway - methodological postings and general questions
of techno-socio-philosophy will tend to be sent to the widest
distribution list.

Sally
----------------------------------------------------------------------
Dr Sally Stowe |Email: stowe-at-rsbs.anu.edu.au
Facility Coordinator |Post: GPO Box 475
ANU Electron Microscopy Unit |ANUEMU (RSBS)
Ph 61 (0)2 6249 2743 |Australian National Univ.
FAX 61 (0)2 6249 4891 |Canberra, Australia 2601
http://online.anu.edu.au/EMU/home.htm







From: MIKE ROCK :      merock-at-du.edu
Date: Mon, 07 Dec 1998 16:04:34 -0700 (MST)
Subject: Re: Immuno

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


BArbara
what antibody are you working with?
how are you processing the tissue?
what are your results from LM of DAB immunostaining?
what dilutions are you currently using with your Abs?
these are a few questions needed to be addressed,
I would be happpy to help once I have more info

-Mike Rock
On Fri, 4 Dec 1998,
Barbara Plowman
wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi Listserver!
} I am doing some immunohistochemistry and have made two attempts but with no
} luck at getting any gold label. My question is whether to use a lower
} dilution gold-antibody or extend the incubation from 4 hours to overnight or
} both? I am a self-confessed rookie at this, so any information on the subject
} would be appreciated .
}
}






From: MIKE ROCK :      merock-at-du.edu
Date: Mon, 07 Dec 1998 16:09:09 -0700 (MST)
Subject: Re: grids dull or shiny?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Elaine-
if you examine grids by SEM you will "see"
there is adifference between the shiny side (which is smoother) than the
dull side of a grid (which is rougher) the rougher side has more surface
area and thus makes a better contact with films or sections, which in turn
keeps them on the grid.
-Mike






From: Paul E. Fischione :      pe_fischione-at-fischione.com
Date: Mon, 7 Dec 1998 18:38:08 -0500
Subject: Available - JEOL 100B TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a JEOL 100B (1974) TEM available at no cost. The microscope is
operational and was converted to operate at 120kV. Also included will be a
second column and HV tank and various other accessories.

Any interested party should contact me as follows:

Paul E. Fischione
E.A. Fischione Instruments, Inc.
9003 Corporate Circle
Export, PA 15632
Phone (724)325-5444
FAX (724)325-5443
e-mail pe_fischione-at-fischione.com







From: marienhoff.visitec-at-t-online.de (Marienhoff)
Date: Mon, 7 Dec 1998 18:37:27 -0500
Subject: motorized table for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,

We want to integrate two separate motorized tables of five axis each into our
large chamber SEM. Who knows suppliers of these tables?

Best regards

Peter Marienhoff
___________________________
Dr. Peter Marienhoff
VisiTec Microtechnik GmbH
Karl-Marx-Str. 14
D-23936 Grevesmuehlen
Germany

Fon: +49-3881-790-47
Fax: +49-3881-790-48
email: pmarienhoff-at-visitec-em.de
http://www.visitec-em.de







From: Alan Bright :      Bright-at-dial.pipex.com
Date: Monday, December 07, 1998 03:43
Subject: Lipstick

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Not a problem a cryostat would make this very easy to section -at- 10 microns
you would need to have specimen temperature control available, and have the
microtome, knife & anti-roll set to 15-20 deg C colder than the lipstick.

Best Regards


Alan Bright

Bright Instrument Co.Ltd.
St Margarets Way
Huntingdon
PE18 6EB
England

Tel No; 01480 454528
Fax No;01480 456031
Email ; Bright-at-dial.pipex.com



-----Original Message-----
} From: Barbara Weyn {ingridb-at-uia.ua.ac.be}
To: Microscopy-at-Sparc5.Microscopy.Com {Microscopy-at-Sparc5.Microscopy.Com}







From: Tim P. LaFave Jr. :      lafatim-at-charlie.cns.iit.edu
Date: Mon, 7 Dec 1998 19:08:51 -0600 (CST)
Subject: RE: New listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Personally, the quantity of information, though somewhat diverse, is
perfectsly fine. I agree, be more clear w.r.t. subject headings. I usually
delete a handful or more of msg's before opening any if they appear to not
be of interest. I also subsc. to about five listservers in other areas,
and run then all through the same email account. Roughly 500-750
emails per day on this account alone. Absolutely no problems.

Leave the Microscopy list alone. Much of the Bio-type information has been
relevant to my work in Materials-type research, etc.

TJ


__ _-==-=_,-.
/--`' \_-at---at-.-- {
Tim (TJ) LaFave Jr. `--'\ \ {___/.
Department of Physics \ \\ " /
University of North Carolina, Charlotte } =\\_/` {
Charlotte, NC 28223 ____ /= | \_|/
_' `\ _/=== \___/
(704)547-3392 [x4] `___/ //\./=/~\====\
(704)509-6622 [Hm] \ // / | ===:
http://www.iit.edu/~lafatim | ._/_,__|_ ==: __
\/ \\ \\`--| / \\
---------- +*+ ---------- | _ \\: /==:-\
`.__' `-____/ |--|==:
Such that the future be theirs \ \ ===\ :==:`-'
to shape and direct. _} \ ===\ /==/
-----------------------------------------------------------------

On Mon, 7 Dec 1998, Marti, Jordi wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I tend to agree with Chuck.
}
} So far the listserver has been a good source of infomation. I think we
} would all lose if we splinter into smaller listservers. If anything I
} would consider using better (i.e. clearer) Subject Identifiers. For
} example, one could use headers such as Bio (for Biological), Mat (for
} Materials ) and Gen. (for general ) subjects. This would make it easier for
} a reader to decide if he/she wants to open a given posting. I believe
} something like this was proposed by others in the past.
}
} We also get much useful information from vendors (I am not a vendor) and
} if we were to split this would pose increased problems for both vendors
} as well as us microscopists in terms of the time spent searching and
} posting messages.
}
} In short, I think we would lose more than what we would gain.
}
} Jordi Marti
}
}
} ----------
} } From: Garber, Charles A.
} To: MICROSCOPY BB
} Subject: New listserver
} Date: Saturday, December 05, 1998 11:49AM
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} -- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --
}
} Dr. Rik Brydson wrote:
} =====================================
} Some information on a new listserver recently set up. I encourage you to
} join Please display in your institution.
}
} Many thanks,
} ================================================
} It is not clear to me how this new listserver would in some way contribute
} to the objectives as outlined, in ways not possible via the existing
} listserver sponsored by the Microscopy Society of America. Right now, if
} someone has a question, they need post it in only one place, if the "places"
} were fragmented, to get the same full coverage, one would have to place it
} in more than one place.
}
} And for those who felt they wanted to subscribe to more than one listserver,
} they are going to be reading the same postings more than one time.
}
} I would cast my vote that the system we have "ain't broken" and therefore we
} shouldn't try to fix it!
}
} Chuck
}
}
}
} ===================================================
} Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
} President 1-(800)-2424-SPI
} SPI SUPPLIES FAX: 1-(610)-436-5755
} PO BOX 656 e-mail: cgarber-at-2spi.com
} West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com
}
}
} Look for us!
} ############################
} WWW: www.2spi.com
} ############################
} ==================================================
}
}






From: Tim P. LaFave Jr. :      lafatim-at-charlie.cns.iit.edu
Date: Mon, 7 Dec 1998 19:14:06 -0600 (CST)
Subject: RE: New listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


.again, a concurrence. Notes from abroad (from the U.S.) are usually the
notes to watch for. :) IN America I've noticed time and again that we want
to 'finitize' everything. Hence the reason I initially avoided Electrical
Engineering as a course of study. Leave the list as it is, perhaps tighten
up the content, subject lines, and make a more strict statement RE:
off-topic posts (LIKE THIS ONE!) --i.e. forbid them.

TJ

__ _-==-=_,-.
/--`' \_-at---at-.-- {
Tim (TJ) LaFave Jr. `--'\ \ {___/.
Department of Physics \ \\ " /
University of North Carolina, Charlotte } =\\_/` {
Charlotte, NC 28223 ____ /= | \_|/
_' `\ _/=== \___/
(704)547-3392 [x4] `___/ //\./=/~\====\
(704)509-6622 [Hm] \ // / | ===:
http://www.iit.edu/~lafatim | ._/_,__|_ ==: __
\/ \\ \\`--| / \\
---------- +*+ ---------- | _ \\: /==:-\
`.__' `-____/ |--|==:
Such that the future be theirs \ \ ===\ :==:`-'
to shape and direct. _} \ ===\ /==/
-----------------------------------------------------------------

On Mon, 7 Dec 1998, Barbara Foster wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Dear all,
}
}
} Usually I try to stay out of the "political" discussions, but I would
} like to put my two cents in on this one.
}
}
} My initial training in microscopy came from the Royal Microscopical
} Society in England. One thing which differentiated their approach from
} more typical US courses was their cross-disciplinary perspective:
} electron microscopists talked to light microscopists and materials
} scientists learned new staining approaches from biologists, etc. We all
} benefit from the diverse postings on the MSA listserver. I agree with
} Jordi, that headers would be helpful when our mailboxes get too full, but
} I, for one, learn as much as possible about "what the other guy is doing"
} so that I can add to my own experience base.
}
}
} By the way: as an avid "industry watcher", I can tell you with certainty
} that the current trend in professional societies is toward partnering
} with sister-disciplines. The fact that our annual meeting is now M&M,
} not just MSA, should be evidence enough. Let's keep the splintering to a
} minimum and use headers, instead, to divvy up the info.
}
}
} Hope this is helpful.
}
} Barbara Foster
}
} Consortium President
}
} {color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
} ..Educating microscopists for greater productivity.
}
}
} {/color} 125 Paridon Street Suite 102 Springfield, MA 01118
}
} PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
}
} Visit our web site { {http://www.MME-Microscopy.com/education}
}
} ******************************************************
}
} {bigger} {bigger} MME {/bigger} {/bigger} is America's first national
} consortium dedicated to
}
} customized on-site training in all areas of
}
} microscopy, sample preparation, and image analysis {bigger} .
}
} {/bigger}
}
}
}
}
}
}
} At 08:10 AM 12/7/98 -0700, Marti, Jordi wrote:
}
} } ------------------------------------------------------------------------
}
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
}
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
}
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
} } -----------------------------------------------------------------------.
}
} }
}
} }
}
} } I tend to agree with Chuck.
}
} }
}
} } So far the listserver has been a good source of infomation. I think
} we
}
} } would all lose if we splinter into smaller listservers. If anything
} I
}
} } would consider using better (i.e. clearer) Subject Identifiers. For
}
} } example, one could use headers such as Bio (for Biological), Mat (for
}
} } Materials ) and Gen. (for general ) subjects. This would make it easier
} for
}
} } a reader to decide if he/she wants to open a given posting. I
} believe
}
} } something like this was proposed by others in the past.
}
} }
}
} } We also get much useful information from vendors (I am not a vendor)
} and
}
} } if we were to split this would pose increased problems for both
} vendors
}
} } as well as us microscopists in terms of the time spent searching
} and
}
} } posting messages.
}
} }
}
} } In short, I think we would lose more than what we would gain.
}
} }
}
} } Jordi Marti
}
} }
}
} }
}
} } ----------
}
} } } From: Garber, Charles A.
}
} } To: MICROSCOPY BB
}
} } Subject: New listserver
}
} } Date: Saturday, December 05, 1998 11:49AM
}
} }
}
} }
} ------------------------------------------------------------------------
}
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of
} America
}
} } To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
}
} } On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
} }
} -----------------------------------------------------------------------.
}
} }
}
} }
}
} } -- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --
}
} }
}
} } Dr. Rik Brydson wrote:
}
} } =====================================
}
} } Some information on a new listserver recently set up. I encourage you
} to
}
} } join Please display in your institution.
}
} }
}
} } Many thanks,
}
} } ================================================
}
} } It is not clear to me how this new listserver would in some way
} contribute
}
} } to the objectives as outlined, in ways not possible via the existing
}
} } listserver sponsored by the Microscopy Society of America. Right now,
} if
}
} } someone has a question, they need post it in only one place, if the
} "places"
}
} } were fragmented, to get the same full coverage, one would have to place
} it
}
} } in more than one place.
}
} }
}
} } And for those who felt they wanted to subscribe to more than one
} listserver,
}
} } they are going to be reading the same postings more than one time.
}
} }
}
} } I would cast my vote that the system we have "ain't broken" and
} therefore we
}
} } shouldn't try to fix it!
}
} }
}
} } Chuck
}
} }
}
} }
}
} }
}
} } ===================================================
}
} } Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
}
} } President 1-(800)-2424-SPI
}
} } SPI SUPPLIES FAX: 1-(610)-436-5755
}
} } PO BOX 656 e-mail: cgarber-at-2spi.com
}
} } West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com
}
} }
}
} }
}
} } Look for us!
}
} } ############################
}
} } WWW: www.2spi.com
}
} } ############################
}
} } ==================================================
}
} }
}
} }
}
} }
}
} }
}
}
}






From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Mon, 7 Dec 1998 17:16:00 -0800 (PST)
Subject: Thanks & responses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To All Who Responded To My Posting,


Thanks for telling me what was wrong with the address to Microscopy
Today. For those of us who are not real literate with computers the answer
was the fact that there were 2 -at- symbols in the address. From now on I
will check to make sure there is only one -at-. But does anyone know why
there were 2 -at-'s in the address to begin with?

Another comment is to the people to my big sigh...
I've heard from several people who thought I jumped the gun by
commenting before all the data was presented. That is true but it still
does not address the fact the women are paid less than men for the same
work. Now, guys, don't get in a huff, but that has been true for years.

Plus to ease the mind of others who feel underpaid because of what
the averages were (who like me are paid a lot less than the average), we
need to see the total results & find out the demographics of the whole
survey.
I tried to respond to the survey, but my illiteracy with computers &
e-mail prevented me from sending my answers (see above).

Think of it this way, there must be some great paying jobs out
there. We just have to know where to find 'em.


Paula :-)

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu
phone: 510-642-2085
fax: 510-643-6207
http://biology.berkeley.edu/EML







From: South Bay Technology :      Henriks-at-CompuServe.COM
Date: Mon, 7 Dec 1998 20:42:19 -0500
Subject: Magnification calib in STEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dr. Frank:

One solution to your calibration problem is using the MAG*I*CAL TEM
Calibration Standard.

The MAG*I*CAL is a TEM calibration standard that performs all of the thr=
ee
major instrument calibrations for a TEM: image magnification; camera
constant for indexing diffraction patterns; and image/diffraction patter=
n
rotation for relating crystal directions to features in the image. =

MAG*I*CAL consists of an electron transparent cross-sectional TEM sample
made from a MBE grown, single-crystal semiconductor wafer. When the
calibration structure is viewed in a TEM, it appears as a series of light=

and dark layers where the layer thicknesses are accurately known. The
calibrated thickness measurements of these light (silicon) and dark (SiGe=

alloy) layers are based on careful TEM measurements of the {111} lattice=

spacing of silicon which is visible on the calibration sample itself, and=

are supported by x-ray diffraction measurements. The layer spacings are
designed so that the sample can be used to calibrate the entire
magnification range in a TEM - from 1,000X to 1,000,000X. As the sample =
is
also a single crystal of silicon, the calibrations requiring electron
diffraction information such as the camera constant and image/diffraction=

pattern rotation can also be performed easily and unambiguously. One
single calibration sample can therefore be used to provide all three of t=
he
major TEM instrument calibrations at all magnifications and all camera
lengths.

With regard to the traceability and certification of the MAG*I*CAL(TM)
calibration sample, each sample is grown on {001} oriented single crystal=

silicon, and all spacings on the sample are directly referenced to the =

cross-sectional (111) lattice spacing of silicon. This spacing is visibl=
e
by lattice imaging on the sample itself, giving each sample the capabilit=
y
of being self-calibrating. Each unit comes with a numbered certificate,=

the =

text of which is included below. This certificate has been used for ISO
9000 certification, with the argument that to our knowledge, this is the
highest quality TEM sample available anywhere in the world at this time. =
=


The MAG*I*CAL (TM) calibration sample consists of sets of thin, nominally=

10 nm alloy layers of Si0.81Ge0.19 alternating with 10 nm pure silicon
layers, on a single crystal silicon {001} substrate. These electronic
device quality layers were grown by Molecular Beam Epitaxy (MBE) as
strained layers, i.e., the alloy layers have a slightly different crystal=

lattice constant, but are strained to conform to the lattice spacing of
pure silicon, so that the material remains single crystal. Lattice image=
s
should therefore be taken in the region of the sample containing no Ge, b=
ut
other measurements are unaffected. The layer thickness variation across t=
he
wafer was measured by double crystal x-ray diffraction (DCXRD) mapping as=
{
1.0%. =


All four sets of the five thin Si0.81Ge0.19 alloy layers and alternating
pure silicon layers (superlattices) were directly calibrated by high
resolution transmission electron microscopy (HREM) with the cross-section=
al
(111) =

lattice spacing of the single crystal silicon substrate, equal to 0.31354=
3
nm [1]. These measurements are also supported by (DCXRD). =


The error in all spacings in the superlattices is one atomic layer:

=A6t =3D +0.3 nm or approximately +3%
The larger, nominally 1.0 micron silicon spacings were calibrated against=

these superlattices. The total error across the entire calibration sampl=
e
is given as:

=A6t =3D + 3%

[1] CRC Handbook of Chemistry and Physics, CRC Press, Inc., Boca Raton,=

Florida 33431


I also have copies of other research papers that have been written by th=
e
developer, John Mccaffrey,
which will provide you with much greater detail. If you have an interest=
,
please let me know and I'll forward the information to you.

If you require any additional information, please feel free to contact me=




From: Michael Dunlap :      mrdunlap-at-ucdavis.edu
Date: Mon, 7 Dec 1998 17:53:27 -0800
Subject: Z-6040 Where to purchase?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have a student that would like to try to section his material. Where can
he purchase Z-6040?

Thanks

Mike

============================================================
Michael Dunlap lab (530) 752-0284
Facility For Advanced Instrumentation fax (530) 752-4412
9-Hutchison Hall
University of California mrdunlap-at-ucdavis.edu
One Shields Ave.
http://FAI.ucdavis.edu
Davis CA, 95616-8594
============================================================







From: Peter Jordan :      emsi-at-pe.net
Date: Mon, 07 Dec 1998 21:02:30 -0800
Subject: Zeiss 10C

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi:
I am still looking for a Zeiss 10C. Will pay top $$$ for it. Also, if
you need parts for a Philips 300 please let me know.
Peter Jordan, EMSI 909 694-1839





From: ard-at-ansto.gov.au (Arthur Day)
Date: Tue, 8 Dec 1998 17:17:48 +1100
Subject: Re: Salary, Old Boys, and New Listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} there is still an "old boys
} network" in place.

Yes it's nagging social issue, but cripes, maybe we should split into
separate male and female microscopy listservers.........










(just a joke)








From: Bettina :      B.Wolpensinger-at-unsw.edu.au
Date: Tue, 8 Dec 1998 19:12:28 +1100
Subject: salary + negative staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} } there is still an "old boys
} } network" in place.
}
} Yes it's nagging social issue, but cripes, maybe we should split into
} separate male and female microscopy listservers.........
}
}
}
} (just a joke)

I am glad that the sense of humor hasn't totally gone lost...... Thanks Arthur!
Thanks mate!

I have followed the discussion about this hot topic with increasing interest. I
somehow agree with those saying that you should only believe in a statistic
which you have manipulated by yourself.

But even if the survey is not really reliable I am pretty sure that most of us
agree that the result does match the general knowledge about discrimiation and
that also might be the reason for those frustrated comments. I am happy that the
situation of women has increased over the last decades (otherwise I might not be
able to post this contribution here), but it is much too early to lean back
there is still a lot to fight for.

In my opinion the salary question is just one out of an uncountable number of
symptoms, the origin is located much deeper. Just to give some keywords: equal
education, who sacrifices the career for looking after kids... and so on...

I remember a conference annoucement on this listserver which said: "Dear
Sirs...". OF COURSE I complained and finally it came out that it was because of
a misunderstanding and I got a big apologize. Okay. But as long as things like
this happen at all, I know that discrimination is present!!

Now to make this contribution seriously scientific: There was a question for
literature about negative staining recently. Have a look at: "Has negative
staining still a place in biomacromolecuar electron microscopy?", A. Bremer, Ch.
Henn, A. Engel, W. Baumeister, U. Aebi, Ultramocroscopy 46 (1992) 85-111

Bettina



***
Bettina Wolpensinger
Electron Microscope Unit
University of New South Wales
Sydney, NSW 2052
AUSTRALIA
phone: +612 9385 6390
fax: +621 9385 6400
b.wolpensinger-at-unsw.edu.au
http://srv.emunit.unsw.edu.au
***






From: ricardo :      ricardo-at-ans.com.au
Date: Tue, 8 Dec 1998 21:16:42 +1100
Subject: I have seen very nice microscope....

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have seen very nice microscop , called RCH.....

The Relief Contrast according to Hostounsky, RCH makes it possible to
observe objects possessing low natural contrast i.e. possessing almost the
same index of refraction as their surroundings. This new contrast enhancing
method makes use of interaction of skew monochromatic rays with the aperture
of RCH condenser.
This device consists of an Abbe-type condenser (numerical aperture 1.2), a
relief diaphragm, and a tiltable interference filter.
The aperture iris diaphragm of the condenser is controlled by a ring with an
aperture scale 0.1 - 1.3, and the relief diaphragm is operated by a lever on
the condenser side between 0 and I. The contrast enhancement in the position
"O" is minimal and in the position "I" maximal. A knob (also on the
condenser side) changes the interference filter tilt which controls the
wavelength of the monochromatic light. each condenser is provided with a
table showing the wavelength which belong to the scale numbers showing the
tilt size. In the set is also contained a wrench serving for insertion and
removal of the condenser from its bracket in the substage condenser carrier.
The RCH condenser for the relief contrast can by applied only with
microscope objectives with magnification 10 : 1 and more.
Advantage of the Houstounsky's relief Contrast Method is most esteemed by
scientists studying living objects: micro-organisms, cells in tissue
cultures, yeast cells, sperm cells, algae, etc. The device is applicable in
research laboratory and in medical diagnostic laboratory, ecological and
hydrobiological laboratory, and also in school training.

If you need more information do not hesitate to ask.

Price is from US$ 1075 plus postage and handling.
attachment for CCD camera and photo available.

Orders:
LAMBDA PRAHA Inc.
Musilkova 12/448
150 00 Praha 5
Czech Rep.
Phone Fax: (420+2) 572 107 60


Keep care and be of good cheer.

Regards

Vratislav Richard Eugene Maria John Baptiste
of Bejsak (Bayshark)-Collorado-Mansfeld

Coleoptera - Australia, Tenebrionidae of World
(incl. Lagriinae, Alleculinae)

Temporally home address:
32 Girrawheen Ave.
Kiama NSW 2533
Australia
e-mail: vratislav-at-bigfoot.com
ricardo-at-ans.com.au
(before Ricardo-at-compuserve.com
and ricardo-at-login.cz )
phone : 0061 0414 540 465

Only after the last tree has been cut down,
only after the last river has been poisoned,
only after the last fish has been caught,
only then will you find that money can not be eaten.'
CREE INDIAN PROPHECY.









From: Richard Edwards :      REE3-at-leicester.ac.uk
Date: Tue, 8 Dec 1998 10:32:35 +0100 (BST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




unsubscribe






From: ricardo :      ricardo-at-ans.com.au
Date: Tue, 8 Dec 1998 22:03:02 +1100
Subject: Re: New listserver

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear colleagues,

I deeply agree with Chuck.

To split listserver into small section do not work, or all section will
slowly dissapear. I can see it on Entomology listserver and Buprestidae
listserver, first still work well but Buprestidae listserver has zero
traffic in this time...

Keep care and be of good cheer.

Regards

Vratislav Richard Eugene Maria John Baptiste
of Bejsak (Bayshark)-Collorado-Mansfeld

Coleoptera - Australia, Tenebrionidae of World
(incl. Lagriinae, Alleculinae)

Temporally home address:
32 Girrawheen Ave.
Kiama NSW 2533
Australia
e-mail: vratislav-at-bigfoot.com
ricardo-at-ans.com.au
(before Ricardo-at-compuserve.com
and ricardo-at-login.cz )
phone : 0061 0414 540 465

Only after the last tree has been cut down,
only after the last river has been poisoned,
only after the last fish has been caught,
only then will you find that money can not be eaten.'
CREE INDIAN PROPHECY.








From: Jonathan Barnard :      J.Barnard-at-bristol.ac.uk
Date: Tue, 08 Dec 1998 11:12:07 +0000
Subject: Re:STEM calibration - error in last message.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I said previously that you should use the (0001) orientation- which is
wrong -as this is down the c axis.
The orientation of the 6H - SiC you should use is somewhere between
[1-100] and the [11-20] directions i.e. with the (000n) planes
perpendicular to the beam.
Sorry for the mistake.

Jon

--
*****************************************
Jonathan Barnard

Microstructural Physics,
H.H.Wills Physics Laboratory,
University of Bristol,
Tyndall Avenue,
Bristol BS8 1TL.

0117 928 9000 ext 8750

*****************************************







From: Yvan Lindekens :      yvan.lindekens-at-skynet.be
Date: Tue, 8 Dec 1998 11:42:28 +0100
Subject: Re: LM Glycerol Borate mount medium?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Scott,

Perhaps you mean glycerin jelly with borax acc. to Fischer? The original
recipe was published in
Ztschr. f. wiss. Mikr., XXIX, p. 65, 1912.

The main advantage of this solution is, that it is liquid at RT... It's
said that it hardens rapidly.

This is the recipe as quoted in Langeron, M.: "Precis de microscopie",
1934, p. 603.

Distilled water 240 g
borax (= Na2B4O7): 5g
glycerin: 25g
gelatin: 40g

Dissolve all at RT. After solution put "some time" in a paraffin incubator.
This solution needs to be neutral...

I don't know the Ri of this medium...

Have you considered von Apathy's medium for your purpose?

Hope this helps...

Yvan Lindekens, Belgium.


} The little info I have suggests that Glycerol Borate might be a nearly
ideal
} mounting medium for my purposes. But more info seems hard to find.
} Scott Harden






From: Miller.David :      miller-at-niehs.nih.gov
Date: Tue, 8 Dec 1998 07:58:44 -0500
Subject: Postdoctoral Position Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


POSTDOCTORAL POSITION - CELL PHYSIOLOGY
I have an opening for a Postdoctoral Fellow in my laboratory . The research
is concerned with the cellular and molecular biology of drug transport
across the renal tubule and the blood-brain barrier. We use isolated renal
proximal tubules, renal cells in culture, isolated brain capillaries,
fluorescent substrates and confocal microscopy to follow transport across
cells. Our goals are to 1) characterize the fundamental membrane-based and
intracellular processes involved, 2) determine how they are controlled by
hormones and xenobioitcs, and 3) identify the signal transduction pathways
involved. Candidates should have a Ph.D. or M.D. degree, less than 5 years
of postdoctoral experience and a background in cellular physiology, membrane
transport or renal physiology. This is a non-tenure track position (NIH IRTA
Fellow).







From: Susan Carbyn :      CARBYNS-at-em.agr.ca
Date: Tue, 08 Dec 1998 08:10:13 -0500
Subject: Automatic Specimen processor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

A few months ago, there was a thread on Automatic Specimen processors. At that time, we were not interested in purchasing one, but certain circumstances have led us to reconsider. I need some prices on them, and some feed back from anyone who may have purchased one and could provide me with some information on the advantages and/or disadvantages with their product. If people could reply to me directly, I would greatly appreciate. I need this information ASAP (like today), so I can put forth the request to the powers that be.

Quotations and information can also be faxed to the number below.

Cheers,

Susan

Susan Carbyn (Electron Microscopist)
Atlantic Food and Horticulture Research Centre
Agriculture and Agri-Food Canada
Kentville, Nova Scotia B4N 1J5
Canada

Phone (902) 679-5566
Fax (902) 679-2311

E-mail: carbyns-at-em.agr.ca
!
!






From: Nancy Zjaba :      Nancy.Zjaba-at-nsc.com
Date: 08 Dec 1998 05:14:41 -0800
Subject: Re: salary survey thoughts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


---------------------- Forwarded by Nancy Zjaba/Americas/NSC on 12/08/98 08:17
AM ---------------------------


Nancy Zjaba
12/08/98 07:29 AM
To: B.Wolpensinger-at-unsw.edu.au -at- Internet
cc:

Interesting thread on the salary survey results...

I've worked in various settings, not all scientific, and one of the things I
have found interesting is the reaction of some men to blatant discrimination
against the women they work with. While they have not been directly
responsible for the discrimination, they simply sit back and observe silently.
They are in no way inclined to alter a situation that benefits them.
Logically, why should they?

I think that's a big part of the issue. For discrimination of any kind to end,
the members of the favored group must exercise some form of moral conviction to
stop the discrimination. Sometimes that happens, sometimes it doesn't.

Nancy Zjaba
National Semiconductor (a very good place to work, I must add)
South Portland, ME







From: Scott Ireland :      sireland-at-frontiernet.net
Date: Tue, 8 Dec 1998 08:16:49 -0500
Subject: the old boys network

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by node21.frontiernet.net (8.8.8a/8.8.8) with SMTP id IAA94146;
Tue, 8 Dec 1998 08:25:06 -0500


Let me put forth a 'possible' (i.e., no data to support this, just
observation) explanation - as a vendor who has worked with the microscopy
community for a few years, I have a noticed that many more women are
entering this profession than men. I believe that I have read that this is
true in most science (not necessarily engineering) fields today. If that
observation is correct, then perhaps the lower average salary is a result of
greater numbers of women new to the field.

Yes, there does seem to be an old boys network at the top, but given the
sheer numbers of women entering the field, that can't stay like that
forever. I don't want to belittle this if it truly is a case of women doing
the same job and being paid less - that certainly is not right - but I think
that if my observations are correct, and have contributed to this sampling
error - then maybe in the long run it is a good sign, and as these new
microscopists gain experience the median salary will rise, and the
differential will decrease.

Just a thought, and only my own - not my employers -


------------------------------
Scott Ireland
North American Sales Manager
Media Cybernetics, L.P.
"The Imaging Experts"
716.473.0222 Tel
716.473.8048 Fax
888.691.2492 Pager
scott-at-mediacy.com
http://www.mediacy.com
http://www.optimas.com
-----------------------------

-----Original Message-----
} From: Dr. Rachel Teitelbaum [mailto:teitelba-at-aecom.yu.edu]
Sent: Monday, December 07, 1998 5:26 PM
To: Douglas W. Darnowski
Cc: microscopy-at-sparc5.microscopy.com


So how does one explain the dominance of men in the field, as evidenced by
fewer tenured professors, etc., and I don't know if NIH breaks down its
grants this way, but there are certainly many fewer Howard Hughes
recipients who are female, and nobel laureates for that matter. I think
it is a rational deduction that there is much less respect for women in
science, and that despite the changing times, there is still an "old boys
network" in place. Surveys in the business field have certainly
demonstrated that women are often paid less despite similar
qualifications. I'm sorry that more people did not respond to the survey
to achieve statistical significance, as to how the results might have
changed with more responses.....

-Rachel








From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Tue, 08 Dec 1998 08:43:58 -0500
Subject: A request of information /polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


maybe some of you guys could help this gentleman out. This is out of my
field here. Reply to him directly as he is not yet a member of this list.
His email address is:

encosrl-at-tin.it


thanks







} From: "ENCO Srl" {encosrl-at-tin.it}
} To: {sdw-at-biotech.ufl.edu}
} Subject: A request of information
} Date: Tue, 8 Dec 1998 11:50:30 +0100
} X-MSMail-Priority: Normal
} X-Mailer: Microsoft Outlook Express 4.71.1712.3
} X-MimeOLE: Produced By Microsoft MimeOLE V4.71.1712.3
}
} Dear Ladies and Gentlemen I am an Italian student of physics at
} university of Padua (Italy). I am going to work on a thesis about wetting
} and adhesion properties of liquid on glass and polymers . I am very
} interested on finding correlations between the surface energetical
} properties of glass and polymers surfaces determinated by macroscopic
} wettability and contact angle measurements and their microscopic structure
} studied by spectroscopic technique. At the moment, one argument could be
} the problems arising when pharmaceutical products become in contact with
} glass or polymeric compounds; some pharmaceutical products in fact can
} adhere on glasses or polymers loosing or chancing in that way their
} properties. Unfortunately I do not know exactly which technique I have
} better use and I can not find any good references about the idea to
} correlate wettability to those microscopic technique. On your well
} organised web site I could understand you are dealing with similar
} problems everyday. Would you mind helping me, please ? I need to know if
} it is possible: Where is possible to find some bibliographic
} references If someone has already studied the problem of
} wettability of glasses and polymers by liquids How to correlate the
} microscopic structure and/or chemical composition of glass surfaces to
} contact angle measurement. Indeed, please I would like to know what
} do you think about these project: composition can changed and so
} contact angle and wettability measurement can change . ? ) and
} then correlate the changes of contact angle and wettability to different
} chemical structure of glasses surfaces . Please may you tell me what you
} sincerly think about this project ? My adress is: encosrl-at-tin.it Best
} regard and of course thank you very much . Via filande, 13 30038
} Spinea (Venice), Italy.




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {
GO GATORS
Scott D. Whittaker 218 Carr Hall
EM Technician Gainesville, FL 32610
University Of Florida ph 352-392-1184
ICBR EM Core Lab fax 352-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/
The home of " Tips & Tricks "










From: zaluzec-at-Sparc5.Microscopy.Com
Date: Tue, 8 Dec 1998 07:56:19 -0500
Subject: A request of information /polymers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



maybe some of you guys could help this gentleman out. This is out of my
field here. Reply to him directly as he is not yet a member of this list.

His email address is:

encosrl-at-tin.it



} From: "ENCO Srl" {encosrl-at-tin.it}
} To: {sdw-at-biotech.ufl.edu}
} Subject: A request of information
} Date: Tue, 8 Dec 1998 11:50:30 +0100
} X-MSMail-Priority: Normal
} X-Mailer: Microsoft Outlook Express 4.71.1712.3
} X-MimeOLE: Produced By Microsoft MimeOLE V4.71.1712.3
}
} Dear Ladies and Gentlemen I am an Italian student of physics at
} university of Padua (Italy). I am going to work on a thesis about wetting
} and adhesion properties of liquid on glass and polymers . I am very
} interested on finding correlations between the surface energetical
} properties of glass and polymers surfaces determinated by macroscopic
} wettability and contact angle measurements and their microscopic structure
} studied by spectroscopic technique. At the moment, one argument could be
} the problems arising when pharmaceutical products become in contact with
} glass or polymeric compounds; some pharmaceutical products in fact can
} adhere on glasses or polymers loosing or chancing in that way their
} properties. Unfortunately I do not know exactly which technique I have
} better use and I can not find any good references about the idea to
} correlate wettability to those microscopic technique. On your well
} organised web site I could understand you are dealing with similar
} problems everyday. Would you mind helping me, please ? I need to know if
} it is possible: Where is possible to find some bibliographic
} references If someone has already studied the problem of
} wettability of glasses and polymers by liquids How to correlate the
} microscopic structure and/or chemical composition of glass surfaces to
} contact angle measurement. Indeed, please I would like to know what
} do you think about these project: composition can changed and so
} contact angle and wettability measurement can change . ? ) and
} then correlate the changes of contact angle and wettability to different
} chemical structure of glasses surfaces . Please may you tell me what you
} sincerly think about this project ? My adress is: encosrl-at-tin.it Best
} regard and of course thank you very much . Via filande, 13 30038
} Spinea (Venice), Italy.









From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Tue, 8 Dec 1998 14:17:54 +0000 (BST)
Subject: Re: Lipstick

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Cool lipstick to -20C and cut sections in a cryomicrotome using a good
steel knofe.

Patricik Echlin
Cambridge

On Mon, 7 Dec 1998, Barbara Weyn wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} We have to make 10 u sections of lipstick.
} Can somebody help us?
}
}
}
}






From: CMontana4-at-aol.com
Date: Tue, 8 Dec 1998 08:14:47 -0500
Subject: Re: Bio-rad MRC600 & Y2K

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



John,
When I was testing for Y2K conditions in our labs, I found that many
manufactures had not yet tested their own equipment - usually our tests and
results prompted vendor action! We each hold the responsibility of performing
tests on our tools to ensure that, even when the vendor claims compliance,
nothing has been missed. I have developed some basic testing protocols that
can be used on PC based equipment (a booklet for $15). I'm a firm believer
that systems should be tested both before and after a "Y2K fix" has been put
into place. I have seen instances where companies can claim compliance, but
when tested, a system can still fail. While I'm not a software guru, and do
not claim to have the "fix", I've strongly felt a need to improve the general
understanding of the Y2K problem and hope to bring some knowledge and
responsibility to the public and industry concerning this serious problem.
(Sorry to soapbox).

Lisa Montanaro
Consultant
Microscopy/Microscopy Education
ph: (703) 257-7157
fax: (703) 365-2427
e-mail: cmontana4-at-aol.com

In a message dated 12/8/98 12:56:40 AM Mid-Atlantic Standard Time,
jgoodhouse-at-molbio.princeton.edu writes:


{ { Dear Confocal users,
Below is an email I sent to Bio-rad regarding MRC600's and
Y2K. I have received no response from them regarding these issues. Does
anyone out there know the answers to these issues?
Dear Bio-rad,
I am the Confocal / Electron Microscopy Core Facility Manager
for the Department of Molecular Biology at Princeton University. I am
inquiring as to the Y2K status of our MRC600 Confocal Microscope System. We
have had this instrument since 1991, and it continues to be a valuable
instrument for this laboratory, and my
department. We have collected and archived over 150 gigabytes of image
data on optical disks with this instrument. This data needs to be
retrievable currently, as well as in the future. The instrument is
under service contract We are currently running the Comos7.0 program under
Windows 95 software. We are
also running macro programs out of SOM under COMOS version 6.03. These
macro programs are an essential tool for several of my investigators
imaging GFP (green fluorescent protein) in living organisms. Without
these macro programs to run the instrument, we would be unable to obtain
the information required for these experiments. We also run CAS 2.5 and
CAS 3.10 for converting images to tiff formats and for creating 24 bit
merged color tiff files for export to other programs.
I need to know
a) if this instrument will continue operating on and after
01-01-2000,
b) whether or not previously collected images will be read by
the program,
c) if we will continue to be able to run macros under the SOM
command shell,
d) what software and / or hardware upgrades will be required,
and when you will have them available?

Sincerely,

Info & Images at http://www.molbio.princeton.edu/confocal/

Joe Goodhouse
Confocal / EM Core Lab Manager
Department of Molecular Biology
Princeton University
jgoodhose-at-molbio.princeton.edu
609-258-5432
} }







From: Ronald LHerault :      lherault-at-bu.edu
Date: Tue, 8 Dec 1998 09:17:09 -0500 (EST)
Subject: Epon

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since I deal withSEMs, I don't always pay close attention to the LM
topics. It would seem, once again that I have been short sighted.
Another group in our department needs to remove Epon form a sampe to try
decalcifying and reembeding in wax. Is there a way to do this? I thought
there was a method mentioned recently.

TIA

Ron
lherault-at-bu.edu







From: John Shields :      jpshield-at-arches.uga.edu
Date: Tue, 8 Dec 1998 09:28:44 -0500 (Eastern Standard Time)
Subject: Cross Training

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



On Mon, 07 Dec 1998 13:38:17 -0500 Barbara Foster {mme-at-map.com} wrote:
} Usually I try to stay out of the "political" discussions, but I would
} like to put my two cents in on this one.
} My initial training in microscopy came from the Royal Microscopical
} Society in England. One thing which differentiated their approach from
} more typical US courses was their cross-disciplinary perspective:
} electron microscopists talked to light microscopists and materials
} scientists learned new staining approaches from biologists, etc. We all

Thanks for that stereotypic view of American Educational Styles. I'm
glad you at least interjected "typical" in your comment. I can rest
easy now knowing that our center is atypical in its approach to
training microscopists here in Georgia.
********************************************
John P. Shields
Center for Ultrastructural Research
Barrow Hall
University of Georgia
Athens, GA 30602
(706)542-4080
jpshield-at-arches.uga.edu
********************************************






From: Brad Storey :      storey-at-lanl.gov
Date: Tue, 8 Dec 1998 08:00:51 -0700
Subject: Staff Position in Microprobe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


--============_-1298993243==_ma============
Content-Type: text/plain; charset="us-ascii"

Technical Staff Position in Materials Characterization/Microprobe Analysis
Los Alamos National Lab
Nuclear Materials Technology Division
Materials Science and Processing Group (NMT-11)

The Materials Science and Processing Group (NMT-11) of the Nuclear
Materials Technology (NMT) Division at Los Alamos National Laboratory is
seeking a highly motivated scientist interested in the study of plutonium
and other actinide materials. The successful candidate will participate in
ongoing experiments in the Pit Surveillance, Pit Rebuild, Pit
Manufacturing, and MOX Fabrication Programs. The primary experimental
focus will be in maintaining and advancing a microprobe and SEM lab to
perform failure and investigative analysis on production problems and
component failures of actinide bearing samples. The successful candidate
will be expected to design, implement, and perform investigative studies to
better understand the complex behavior of plutonium metal and alloys in the
thermo-mechanically processed, annealed, and aged conditions. The
successful candidate will also be required to perform hands-on work with
plutonium and other actinides in a glovebox environment. Opportunities
exist to participate in other ongoing group and division research.

Required Skills:
Interested applicants shall have a strong materials science background with
demonstrated experience in electron-beam instrumentation and associated
detectors used for materials characterization, e.g., SEM, microprobe, EDS,
WDS, TEM, etc. The position requires strong knowledge and experience in
the area of phase transformations and identification. Effective oral,
written and interpersonal skills are required. The ability to work on
multiple programs concurrently is also required.

Desired Skills:
Demonstrated work experience in an industrial environment is highly
desirable. Demonstrated experience in the handling and processing of
radioactive materials in a glovebox environment.

Education:
A Ph.D. in metallurgy, materials science, or other physical sciences is
required.

Additional Requirements:
This position is subject to the requirements of the Personnel Security
Assurance Program (PSAP). Candidates invited for interview for this
position will be subject to a pre-employment screening check, medical
examination and drug test, and must consent to be in the PSAP program at
the time of the interview. A willingness to work on materials problems
relevant to the safety of the US nuclear weapons stockpile, and the ability
to obtain a "Q" clearance which requires US citizenship are required.

For consideration, submit: resume, publications list, cover letter
outlining current research interests, the names, addresses, and phone
numbers of three references, and one copy of university transcripts to:

Brad G. Storey
Los Alamos National Lab
P.O. Box 1663, MS: E505
Los Alamos, NM 87545
storey-at-lanl.gov
505-667-0458 phone
505-665-4394 fax
Brad G. Storey
Materials Science and Processing Group (NMT-11)
Mail Stop: E505
Los Alamos National Lab
Los Alamos, NM 87545
storey-at-lanl.gov
505-667-0458 phone
505-665-4394 fax
505-996-3129 pager

--============_-1298993243==_ma============
Content-Type: text/enriched; charset="us-ascii"

{bold} {fontfamily} {param} Times {/param} {smaller} Technical Staff Position
in Materials Characterization/Microprobe Analysis

Los Alamos National Lab

Nuclear Materials Technology Division

Materials Science and Processing Group (NMT-11)


{/smaller} {/fontfamily} {/bold} {fontfamily} {param} Times {/param} {smaller} The
Materials Science and Processing Group (NMT-11) of the Nuclear
Materials Technology (NMT) Division at Los Alamos National Laboratory
is seeking a highly motivated scientist interested in the study of
plutonium and other actinide materials. The successful candidate will
participate in ongoing experiments in the Pit Surveillance, Pit
Rebuild, Pit Manufacturing, and MOX Fabrication Programs. The primary
experimental focus will be in maintaining and advancing a microprobe
and SEM lab to perform failure and investigative analysis on production
problems and component failures of actinide bearing samples. The
successful candidate will be expected to design, implement, and perform
investigative studies to better understand the complex behavior of
plutonium metal and alloys in the thermo-mechanically processed,
annealed, and aged conditions. The successful candidate will also be
required to perform hands-on work with plutonium and other actinides in
a glovebox environment. Opportunities exist to participate in other
ongoing group and division research.


{bold} {underline} Required Skills:

{/underline} {/bold} Interested applicants shall have a strong materials
science background with demonstrated experience in electron-beam
instrumentation and associated detectors used for materials
characterization, e.g., SEM, microprobe, EDS, WDS, TEM, etc. The
position requires strong knowledge and experience in the area of phase
transformations and identification. Effective oral, written and
interpersonal skills are required. The ability to work on multiple
programs concurrently is also required.


{bold} {underline} Desired Skills:

{/underline} {/bold} Demonstrated work experience in an industrial
environment is highly desirable. Demonstrated experience in the
handling and processing of radioactive materials in a glovebox
environment.


{bold} {underline} Education:

{/underline} {/bold} A Ph.D. in metallurgy, materials science, or other
physical sciences is required.


{bold} {underline} Additional Requirements:

{/underline} {/bold} This position is subject to the requirements of the
Personnel Security Assurance Program (PSAP). Candidates invited for
interview for this position will be subject to a pre-employment
screening check, medical examination and drug test, and must consent to
be in the PSAP program at the time of the interview. A willingness to
work on materials problems relevant to the safety of the US nuclear
weapons stockpile, and the ability to obtain a "Q" clearance which
requires US citizenship are required.


For consideration, submit: resume, publications list, cover letter
outlining current research interests, the names, addresses, and phone
numbers of three references, and one copy of university transcripts to:


Brad G. Storey

Los Alamos National Lab

P.O. Box 1663, MS: E505

Los Alamos, NM 87545

{underline} {color} {param} 0000,0000,00FF {/param} storey-at-lanl.gov

{/color} {/underline} 505-667-0458 phone

505-665-4394 fax {/smaller} {/fontfamily}

Brad G. Storey

Materials Science and Processing Group (NMT-11)

Mail Stop: E505

Los Alamos National Lab

Los Alamos, NM 87545

storey-at-lanl.gov

505-667-0458 phone

505-665-4394 fax

505-996-3129 pager

--============_-1298993243==_ma============--





From: Douglas W. Darnowski :      darnowsk-at-staff.uiuc.edu
Date: Tue, 8 Dec 1998 09:39:18 -0500
Subject: salary survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rachel, and others watching this thread: (this is my last message to the
thread, if anyone wants to discuss it further, email personally)

} So how does one explain the dominance of men in the field, as evidenced by
} fewer tenured professors, etc., and I don't know if NIH breaks down its
} grants this way, but there are certainly many fewer Howard Hughes
} recipients who are female, and nobel laureates for that matter.

No. Since tenure takes some time to achieve, with a PhD being the usual
first step, and since there has been a change in the demographics, with
regard to gender, it stands to reason that the breakdown of tenure
according to sexual lines may change (indeed, it is changing, as
demographics show). Other factors may intervene, such as time taken away
from work to raise children, which time detracts from effort towards
research and tenure.

As for Howard Hughes and Nobel, consider that those awards are given for
doing certain kinds of research and for certain kinds of acheivements.
Beyond the factors in the paragraph above, consider that those types of
research may not appeal to as many women for some valid reasons. Thus
again, an explanation without need to invoke a bogeyman such as "the old
boys network."

I will add that I have a BS from Yale, a PhD from Cornell, and am a PostDoc
at the U of IL. I am a white man. In the six labs in which I have worked, I
have never reported directly to a white man, and in only two cases was a
white man the PI (one other lab had a white man as PI, but I was doing
some experiments there; he was the only white man on my thesis committee).
I have supervised approximately ten undergraduate
students, of whom none were white non-hispanic males (there were at other
times white male undergrads who I helped with experiments, but none ever
reported to me directly). In
spite of this, I won an NSF, and have consistently received excellent
evaluations and above-average raises, and I have never had a complaint,
directly or indirectly, that I had unfairly treated any of these students.
My personal experience tells me that you don't have to have people with the
same skin color and complement of sex chromosomes to interact with in order
to achieve.

} I think
} it is a rational deduction that there is much less respect for women in
} science, and that despite the changing times, there is still an "old boys
} network" in place.

Respect cannot be deduced from salaries alone. It is a subjective factor,
and you provide no rational evidence for a connection.

} Surveys in the business field have certainly
} demonstrated that women are often paid less despite similar
} qualifications.

Not true. Those surveys have generally been shown to be seriously flawed,
when years on the job and other relevant factors are considered beyond mere
academic qualifications. This topic has been dealt with in the Wall Street
Journal in recent months.

} I'm sorry that more people did not respond to the survey
} to achieve statistical significance, as to how the results might have
} changed with more responses.....

Until you have those responses, it is meaningless to discuss or consider,
which is why those results shouldn't have been posted.

I will add further that my experience has been that such complaints as were
raised by you and Paula are usually made in an emotional way without strong
rational bases (see my criticisms above). Until someone can provide
something beyond assertions of a conspiracy, there is no reason to believe
in them, as far as salaries go. It would be illogical and counter to the
principles on which this nation is founded.

Doug Darnowski



******************************************************************************
Douglas Darnowski
Department of Crop Sciences
384 ERML
1201 West Gregory Drive
University of Illinois
Urbana IL 61801
work ph: (217) 244-6150
fx: (217) 333-4777
home ph: (217) 356-6606
fx: (217) 356-4454
email: darnowsk-at-staff.uiuc.edu
******************************************************************************
Douglas Darnowski
Department of Crop Sciences
384 ERML
1201 West Gregory Drive
University of Illinois
Urbana IL 61801
work ph: (217) 244-6150
fx: (217) 333-4777
home ph: (217) 356-6606
fx: (217) 356-4454
email: darnowsk-at-staff.uiuc.edu





From: Sobocinski, Gregg :      Gregg.Sobocinski-at-WL.com
Date: Tue, 8 Dec 1998 11:31:52 -0500
Subject: Salaries and Discrimination?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thank-you, Rachel T. and Scott I., for responding to his issue in a manner
that appears to be rational and not emotionally generated. I detest the
short, inflaming responses to this list that state emotional responses but
serve no useful purpose, even when it comes to sharing an opinion for the
purpose of debating.

Regarding the "old boy network":
I recently had a friend in the field of archaeology reveal to me that her
married friend within her graduate program was the only one denied financial
support (and teaching experience) because she is married (no children). It
was assumed by influential MALE AND FEMALE advisors in the program that she
had less career potential because she might decide to have children before
she has established herself in her field professionally. This "old boy
network" concept is MORE COMPLEX than just blaming males! Please don't make
me feel defensive or guilty by implying so!

I think this string of responses is a waste of our time unless we get some
more tangible issues to discuss which MAY TAKE US IN A POSITIVE DIRECTION!
As a microscopist without a graduate degree, but a good-paying job in the
field, I was surprised that the "average" salary in the field was so high!
I'm SURE that graduate degrees have a large influence upon this statistic.
(Especially since we have already discussed whether we should be insulted by
some of the job postings to the list looking for experienced microscopists
who will work for $20,000 salaries.)

Note: Along this line of graduate degrees, I know a LOT more females my age
(30ish) who are earning, or have received graduate degrees. I suspect that
in 20 years, the salaries will very likely swing to the advantage of females
(on average). What was the male/female graduate degree ratio 20 years ago
from today?

Thank-you for listening to my longish rant. I hope some productive thoughts
result from it.

Gregg Sobocinski
Parke-Davis Pharmaceuticals
Ann Arbor, Michigan
USA

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Let me put forth a 'possible' (i.e., no data to support this, just
observation) explanation - as a vendor who has worked with the microscopy
community for a few years, I have a noticed that many more women are
entering this profession than men. I believe that I have read that this is
true in most science (not necessarily engineering) fields today. If that
observation is correct, then perhaps the lower average salary is a result of
greater numbers of women new to the field.

Yes, there does seem to be an old boys network at the top, but given the
sheer numbers of women entering the field, that can't stay like that
forever. I don't want to belittle this if it truly is a case of women doing
the same job and being paid less - that certainly is not right - but I think
that if my observations are correct, and have contributed to this sampling
error - then maybe in the long run it is a good sign, and as these new
microscopists gain experience the median salary will rise, and the
differential will decrease.

Just a thought, and only my own - not my employers -


------------------------------
Scott Ireland
North American Sales Manager
Media Cybernetics, L.P.
"The Imaging Experts"
716.473.0222 Tel
716.473.8048 Fax
888.691.2492 Pager
scott-at-mediacy.com
http://www.mediacy.com
http://www.optimas.com
-----------------------------

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} So how does one explain the dominance of men in the field, as evidenced by
} fewer tenured professors, etc., and I don't know if NIH breaks down its
} grants this way, but there are certainly many fewer Howard Hughes
} recipients who are female, and nobel laureates for that matter. I think
} it is a rational deduction that there is much less respect for women in
} science, and that despite the changing times, there is still an "old boys
} network" in place. Surveys in the business field have certainly
} demonstrated that women are often paid less despite similar
} qualifications. I'm sorry that more people did not respond to the survey
} to achieve statistical significance, as to how the results might have
} changed with more responses.....
}
} -Rachel
}





From: Chris Gilpin :      cgilpin-at-fs1.sem.man.ac.uk
Date: Tue, 8 Dec 1998 17:01:43 -0000
Subject: EDX peak significance in quant

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My LINK EDX quant software (ZAFPB) performs a calculation to determine the
significance of the peak quant data. The cut off is 2 Sigma. I have lost my
book (i.e someone has not returned it) which gives the formula for
calculating sigma. Can anyone help out please?
Many thanks
Chris



Chris Gilpin
Experimental Officer
Biological Sciences EM Unit
G452 Stopford Building
Oxford Road
Manchester
M13 9PT
phone +44 0161 275 5170
Fax +44 0161 275 5171 Chris Gilpin
http://www.empgu.man.ac.uk






From: Susan Carbyn :      CARBYNS-at-em.agr.ca
Date: Tue, 08 Dec 1998 12:09:17 -0500
Subject: TEM tissue processor

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I should have made it more clear on what it is I was looking for information about. It is a TEM tissue processor and I apologize to those I confused.

Susan


Susan Carbyn (Electron Microscopist)
Atlantic Food and Horticulture Research Centre
Agriculture and Agri-Food Canada
Kentville, Nova Scotia B4N 1J5
Canada

Phone (902) 679-5566
Fax (902) 679-2311

E-mail: carbyns-at-em.agr.ca






From: Tracey M. Pepper :      tpepper-at-iastate.edu
Date: Tue, 08 Dec 1998 11:11:38 -0600
Subject: my 2 cents

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To the list:
Think about all that's been said.
We are all making good money. Some have more experience, some have less.
Some live in towns where the rent goes for $450 a month for a two bed apt.
some $1200.
We will all put food on our tables, and some of us will soon put presents
in the hands of our friends and family.
We have hope.
We discover.
We have so much more than we "deserve".
Let's be more thankful for what we have, and share a little with those who
don't- regardless of the time of year.


Tracey Pepper
Bessey Microscopy Facility
Iowa State University
ph: 515-294-3872
fax: 515.294.1337






From: Marisa Ahmad :      mahmad-at-semiconductor.com
Date: Tue, 8 Dec 1998 13:29:07 -0500
Subject: W etches

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Anyone out there know of a nice uniform tungsten etch, preferably that will
stop on TiN/Ti or SiO2???? Dry or wet would be fine...any suggestions will
be much much much appreciated.

Thanks in advance,

Marisa Ahmad : )





From: CMontana4-at-aol.com
Date: Tue, 8 Dec 1998 13:26:12 EST
Subject: Re: Bio-rad MRC600 & Y2K

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John,
When I was testing for Y2K conditions in our labs, I found that many
manufactures had not yet tested their own equipment - usually our tests and
results prompted vendor action! We each hold the responsibility of performing
tests on our tools to ensure that, even when the vendor claims compliance,
nothing has been missed. I have developed some basic testing protocols that
can be used on PC based equipment (a booklet for $15). I'm a firm believer
that systems should be tested both before and after a "Y2K fix" has been put
into place. I have seen instances where companies can claim compliance, but
when tested, a system can still fail. While I'm not a software guru, and do
not claim to have the "fix", I've strongly felt a need to improve the general
understanding of the Y2K problem and hope to bring some knowledge and
responsibility to the public and industry concerning this serious problem.
(Sorry to soapbox).

Lisa Montanaro
Consultant
Microscopy/Microscopy Education
ph: (703) 257-7157
fax: (703) 365-2427
e-mail: cmontana4-at-aol.com

In a message dated 12/8/98 12:56:40 AM Mid-Atlantic Standard Time,
jgoodhouse-at-molbio.princeton.edu writes:


{ { Dear Confocal users,
Below is an email I sent to Bio-rad regarding MRC600's and
Y2K. I have received no response from them regarding these issues. Does
anyone out there know the answers to these issues?
Dear Bio-rad,
I am the Confocal / Electron Microscopy Core Facility Manager
for the Department of Molecular Biology at Princeton University. I am
inquiring as to the Y2K status of our MRC600 Confocal Microscope System. We
have had this instrument since 1991, and it continues to be a valuable
instrument for this laboratory, and my
department. We have collected and archived over 150 gigabytes of image
data on optical disks with this instrument. This data needs to be
retrievable currently, as well as in the future. The instrument is
under service contract We are currently running the Comos7.0 program under
Windows 95 software. We are
also running macro programs out of SOM under COMOS version 6.03. These
macro programs are an essential tool for several of my investigators
imaging GFP (green fluorescent protein) in living organisms. Without
these macro programs to run the instrument, we would be unable to obtain
the information required for these experiments. We also run CAS 2.5 and
CAS 3.10 for converting images to tiff formats and for creating 24 bit
merged color tiff files for export to other programs.
I need to know
a) if this instrument will continue operating on and after
01-01-2000,
b) whether or not previously collected images will be read by
the program,
c) if we will continue to be able to run macros under the SOM
command shell,
d) what software and / or hardware upgrades will be required,
and when you will have them available?

Sincerely,

Info & Images at http://www.molbio.princeton.edu/confocal/

Joe Goodhouse
Confocal / EM Core Lab Manager
Department of Molecular Biology
Princeton University
jgoodhose-at-molbio.princeton.edu
609-258-5432
} }





From: Lesley Weston :      lesley-at-interchange.ubc.ca
Date: Tue, 8 Dec 1998 10:36:08 -0800 (PST)
Subject: Parts for microtome

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Many thanks to all those who answered my request for a source of parts for an
MT2 microtome. Someone here in Canada is able to help me (thanks, Chris). It's
good to know there are so many people still using the old machines.

Lesley Weston.









From: Steve Barlow :      sbarlow-at-sunstroke.sdsu.edu
Date: Tue, 8 Dec 1998 11:07:31 -0800
Subject: TEM cold stage wanted

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please respond directly to
"David Bruck" {bruck-at-biomail.sjsu.edu}


} I am searching for a Zeiss EM109 or EM10 cold stage. Please contact me if
} you } know where I can get one.

} David Bruck

---------------------------------------------------------------------

Dr. Steven Barlow, Associate Director
EM Facility/Biology Department
San Diego State University
5500 Campanile Drive
San Diego CA 92182-4614
phone: (619)594-4523
fax: (619) 594-8759
email: sbarlow-at-sunstroke.sdsu.edu
website: http://www.sci.sdsu.edu/emfacility/







From: CMontana4-at-aol.com
Date: Tue, 8 Dec 1998 13:57:05 EST
Subject: Re: motorized table for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Peter,
Do you currently have dual stages that you'd like to fit into your system, or
are you looking for someone to design them for you? There are a few companies
you can talk with (the manufacturer of your SEM would be a place to start),
but I've also spoken with Leo Fama, the president of AMT (508) 774-5550, and
Mark Reynolds of the Ernst Fjeld Company (978) 667-1416 ext. 10. I've worked
with both of these companies to design ultra high precision 5-axis SEM stages
for use in the semiconductor industry.

Let me know how you fare!
Lisa Montanaro
Consultant
Microscopy/Microscopy Education
ph: (703) 257-7157
fax:(703) 365-2427
e-mail: cmontana4-at-aol.com

In a message dated 12/8/98 4:47:34 AM Mid-Atlantic Standard Time,
marienhoff.visitec-at-t-online.de writes:

{ { he Microscopy ListServer -- Sponsor: The Microscopy Society of America
To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
-----------------------------------------------------------------------.


Dear all,

We want to integrate two separate motorized tables of five axis each into our
large chamber SEM. Who knows suppliers of these tables?

Best regards

Peter Marienhoff } }





From: CANTINO-at-ORACLE.PNB.UCONN.EDU (MARIE CANTINO)
Date: Tue, 8 Dec 1998 15:30:53 -0400
Subject: Re: salary survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} It would be illogical and counter to the
} principles on which this nation is founded.

} Doug Darnowski

Are you seriously suggesting that the founding fathers believed men and
women should have equal rights in this country?

Marie

Dr. Marie E. Cantino
Dept. of Physiology and Neurobiology, U-131
University of Connecticut
Storrs, CT 06269
Ph: 860-486-3588
Fax: 860-486-1936








From: Sjoplinh-at-aol.com
Date: Tue, 8 Dec 1998 16:53:43 EST
Subject: Re: LM Glycerol borate mount medium?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks for the info, Yvan. I don't think that's the same stuff, but it sounds
interesting anyway. The medium I'm wondering about was described as "glycerin
borate... once available commercially as "Aqua Resin"...". My impression is
that it's either a compound or a simple solution of glycerin, "borate" (+
water?).

Leonard Corwin kindly provided the following:
{ { FWIW, Beilstein, system number 38 or 39, p. 519 of Band I: boric acid
glycerin triester, glassy yellow mass on heating of glycerin with boric
anhydride (1866).} }

I don't know if that's it either, but doesn't sound that hard to make (?) if
so.

I've not heard of "Apathy's medium"; will try to find references.

I'm looking for a water soluble medium for permanent mounts of chitinous
arthropod appendages. The ideal (?) mountant would; 1) Have a Ri of around
1.46; 2) Have low scatter, unlike gum or gelatin media; 3) Be easy to make
(and / or be available commercially). It would also be nice if it was non -
fluorescent and shrink little or not at all as it dries... but I realize I
can't have everything.

Regards,

Scott Harden








From: Tom Phillips :      tphillips-at-biosci.mbp.missouri.edu
Date: Tue, 8 Dec 1998 16:32:51 -0600
Subject: Re: salary survey - Enough Already!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since the salary survey data concerning microscopists is not available at
the moment, nor does it necessarily have the information required to
adequately interpret it in required to microscopists qualifications, can
the political discussion move off line or to some more relevant listserver.
The current discussion seems to have little relevance to microscopy.

Thomas E. Phillips, Ph.D.
Associate Professor of Biological Sciences
Director, Molecular Cytology Core Facility

3 Tucker Hall
Division of Biological Sciences
University of Missouri
Columbia, MO 65211
(573)-882-4712 (voice)
(573)-882-0123 (fax)





From: HILDEGARD CROWLEY :      hcrowley-at-du.edu
Date: Tue, 08 Dec 1998 15:05:14 -0700 (MST)
Subject: Epon surface-grid:dull,shiny!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi,

If TEM grids are investigated by SEM on both sides, it will be found that
one side is much smoother than the other. One side is very rough. If an
epon section is viewed by SEM both sides are found to be rough (Handbook
of Epoxy Resins). Now, let us say you have two pieces of extra fine
sandpaper. One piece of sandpaper is shiny and slick on one side and rough
on the other. You want the best adhesion (no slippage) when putting the
sandpapers on top of one another. How would you do that?

Hildy






From: Karen Wetmore Grycewicz :      karenw-at-ucmp1.berkeley.edu
Date: Tue, 08 Dec 1998 15:11:46 -0800
Subject: Re: my 2 cents

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 11:11 AM 08-12-98 -0600, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I agree, let's be thankful for what we have, and that many organizations
and people have managed to minimize or eliminate any bias related to sex.

Let's also recognize that there is still a problem at some organizations.
I personally know of some egregious cases (at another institution where I
have friends) that were only recently corrected after the women sued.
When compared to male professors with equivalent academic field, rank,
years at rank, etc, the women had better scientific publication records and
higher citation rates and yet had been consistently given lower raises such
that it took salary increases of more than 20% just to bring some of them
even with male counterparts with lower publication and citation rankings.
Fortunately, these women had objective measures they could use to
demonstrate the problem.

What criteria would one use in the field of microscopy to objectively rank
workers? What criteria are useful but somewhat subjective and therefor
potentially influenced by unconscious biases?

Happy Holidays,

Karen
*******************************************************************
Karen L. Wetmore Grycewicz, Ph.D.
Museum of Paleontology
University of California
Berkeley, CA 94720-4780 karenw-at-ucmp1.berkeley.edu
*******************************************************************





From: Sara Miller :      saram-at-acpub.duke.edu
Date: Tue, 8 Dec 1998 18:16:06 -0500 (EST)
Subject: Re: Zeiss 10C

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Peter,

We have a Zeiss 10A that has been under continuous service contract until
this summer. It is in excellent condition, and is still working fine. I
assume the department is willing to part with it since we will eventually
put a Philips 300 in there. If you're interested, I could ask what
they want for it. What 300 parts are you willing to part with and for how
much?

Sara
(address at end)


On Mon, 7 Dec 1998, Peter Jordan wrote:

} Date: Mon, 07 Dec 1998 21:02:30 -0800
} From: Peter Jordan {emsi-at-pe.net}
} To: Microscopy-at-sparc5.microscopy.com
} Subject: Zeiss 10C
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi:
} I am still looking for a Zeiss 10C. Will pay top $$$ for it. Also, if
} you need parts for a Philips 300 please let me know.
} Peter Jordan, EMSI 909 694-1839
}
}

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735






From: Melvyn Dickson :      M.Dickson-at-unsw.edu.au
Date: Wed, 09 Dec 1998 10:42:43 +1100
Subject: Re: LM Glycerol borate mount medium?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Glycerin Borate jelly

Fischer - Zeit. wiss Mikrosk vol 29, 1912 p65

borax 5 grams
dissolve in 240 ml water
add 25 ml glycerine
add 40 grams gelatine,
dissolve with heat and heat until solution slightly thickens
remains fluid at room temperature

Pantin says glycerine jelly has an RI of 1.47

Apathys Gum syrup has an RI of 1.52

Picked gum arabic 50 grams
cane sugar........50 grams
distilled water 50 ml

dissolve over a water bath

add 0.05 grams thymol

said to set quickly as hard as balsam

recipes from "the Microtomist's Vade-mecum" by Bolles Lee 7th edition
1913.....




*****************************************************
Mel Dickson,
Director.
Electron Microscope Unit,
University of New South Wales.
Sydney NSW 2052 Australia

Phone (+612) 9385-6383
Fax (+612) 9385-6400
Website {http://srv.emunit.unsw.edu.au}
*****************************************************





From: Peter Donahoe :      glencadia-at-berk.com
Date: Tue, 08 Dec 1998 20:23:06 -0500
Subject: Salaries, Old Boys, Lists etc etc

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------9C33972ED628439246ED66C0
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

Yikes Mikes!
You mean to say that microscopists make an average of $46,000 for women
and $55,000 for men? I wish I knew of someone, male or female, around
here making anything near that amount. The school of EM at which I
trained in '96-'97 has been closed [ I was in the last class] and a year
later the whole EM lab shut down - government 'downsizing' of the VA
system I suppose, particularly of redundant pathology facilities? (q.v.
past posts on the future of this application).
Perhaps a bit of perspective on the viability of employment in this
field rather than valid but self-congratulatory carping by those of you
still employed would be
welcome.
Gee, I gotta find the Bubba Micro.list and get me one of them jobs. ;-}
Regards to all. Be of good cheer.

Peter Donahoe - will work for microbes
--------------9C33972ED628439246ED66C0
Content-Type: text/x-vcard; charset=us-ascii; name="vcard.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for Peter or Melanie Donahoe
Content-Disposition: attachment; filename="vcard.vcf"

begin: vcard
fn: Peter or Melanie Donahoe
n: Donahoe;Peter or Melanie
email;internet: glencadia-at-berk.com
x-mozilla-cpt: ;0
x-mozilla-html: FALSE
version: 2.1
end: vcard


--------------9C33972ED628439246ED66C0--






From: Pnina Ari-Gur :      arigurp-at-wmich.edu
Date: Tue, 08 Dec 1998 22:05:48 -0500
Subject: College of Engineering -- Dean's Position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------C23593DE2542DC81AE91F209
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

Please see the attachment (announcement of dean search for our college
of engineering).
--
==============================
Pnina Ari-Gur, D.Sc.
Materials Science and Engineering
Western Michigan University
Kalamazoo, MI 49008
(616) 387-3372 FAX: (616) 387-6517
email: pnina.ari-gur-at-wmich.edu
http://www.wmich.edu/cmd/arigur.htm
==============================


--------------C23593DE2542DC81AE91F209
Content-Type: text/html; charset=us-ascii; name="dean.htm"
Content-Transfer-Encoding: 7bit
Content-Disposition: inline; filename="dean.htm"
Content-Base: "http://www.wmich.edu/engineer/dean.htm"

{html}

{head}
{meta http-equiv="Content-Type" content="text/html; charset=iso-8859-1"}
{title} Job Description of Dean of College of Engineering and Applied Sciences, Western
Michigan University {/title}


{meta name="Microsoft Theme" content="ceas ii 011, default"} {meta name="Microsoft Border" content="tlb, default"} {/head}

{body background="_themes/ceas%20ii/wmichback1.jpg" bgcolor="#FFFFFF" text="#5B3A04" link="#800080" vlink="#3366CC" alink="#FF9900"} {!--msnavigation--} {table border="0" cellpadding="0" cellspacing="0" width="100%"} {tr} {td} {!--mstheme--} {font face="times new roman, times new roman, times"}
{div align="center"} {center}

{!--mstheme--} {/font} {table border="0" cellpadding="0" cellspacing="0" width="800" height="14" align="left"}
{tr}
{td width="164" height="14"} {!--mstheme--} {font face="times new roman, times new roman, times"} {div align="left"} {!--mstheme--} {/font} {table border="0" cellpadding="0" cellspacing="0" width="133"}
{tr}
{td width="222" valign="top" align="left"} {!--mstheme--} {font face="times new roman, times new roman, times"} {p align="center"} {img src="_borders/broncos.jpg" alt="Buster Bronco" WIDTH="110" HEIGHT="87"} {!--mstheme--} {/font} {/td}
{/tr}
{/table} {!--mstheme--} {font face="times new roman, times new roman, times"}
{/div} {!--mstheme--} {/font} {/td}
{td width="612" height="14"} {!--mstheme--} {font face="times new roman, times new roman, times"} {p align="center"} {MAP NAME="FrontPageMap"} {AREA SHAPE="RECT" COORDS="50, 80, 62, 92" HREF="http://www.hotmail.com"} {/MAP} {a href="/_derived/dean.htm.map"} {img ismap usemap="#FrontPageMap" border="0" alt="College of Engineering and Applied Sciences" src="_borders/Banner.jpg"} {/a} {img src="_borders/wmubar.jpg" WIDTH="429" HEIGHT="21"} {!--mstheme--} {/font} {/td}
{/tr}
{/table} {!--mstheme--} {font face="times new roman, times new roman, times"}
{/center} {/div}
{!--mstheme--} {/font} {/td} {/tr} {!--msnavigation--} {/table} {!--msnavigation--} {table border="0" cellpadding="0" cellspacing="0" width="100%"} {tr} {td valign="top" width="1%"} {!--mstheme--} {font face="times new roman, times new roman, times"}
{div align="left"}

{!--mstheme--} {/font} {table border="0" cellpadding="0" width="144" height="81" cellspacing="12"}
{tr}
{td align="center" width="469" height="11"} {!--mstheme--} {font face="times new roman, times new roman, times"} {strong} {big} {big} {font color="#008080" face="Bradley Hand ITC"} WMU {/font} {/big} {/big} {/strong} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="11"} {!--mstheme--} {font face="times new roman, times new roman, times"} {strong} {small} {font face="Bradley Hand ITC"} {a href="http://www.wmich.edu"} WMU Home {/a} {/font} {/small} {/strong} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="11"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="http://www.wmich.edu/registrar/index.html"} {font face="Bradley Hand ITC"} {strong} {small} Winter
1999 Schedule {/small} {/strong} {/font} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="11"} {!--mstheme--} {font face="times new roman, times new roman, times"} {font color="#008080" face="Bradley Hand ITC"} {strong} {big} {big} CEAS {/big} {/big} {/strong} {/font} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="11"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="index.htm"} {strong} {small} {font face="Bradley Hand ITC"} CEAS Home {/font} {/small} {/strong} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="11"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="http://www.wmich.edu/engineer/cae/index.htm"} {strong} {small} {font face="Bradley Hand ITC"} CAE Center Home {/font} {/small} {/strong} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="10"} {!--mstheme--} {font face="times new roman, times new roman, times"} {p align="center"} {a href="http://www.wmich.edu/admi/"} {strong} {small} {font face="Bradley Hand ITC"} Admissions {/font} {/small} {/strong} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="http://www.wmich.edu/alumni/"} {small} {strong} {font face="Bradley Hand ITC"} Alumni {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="undergra.htm"} {small} {strong} {font face="Bradley Hand ITC"} Undergraduate Programs {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="1"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="grad.htm"} {small} {strong} {font face="Bradley Hand ITC"} Graduate Programs {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {font face="Bradley Hand ITC"} {a href="phd.htm"} {small} {strong} Ph.D. Programs {/strong} {/small} {/a} {/font} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="department.htm"} {small} {strong} {font face="Bradley Hand ITC"} Departments {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="faculty.htm"} {small} {strong} {font face="Bradley Hand ITC"} Faculty Index {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="research.htm"} {small} {strong} {font face="Bradley Hand ITC"} Research and Testing {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="http://www.wmich.edu/engineer/Schship/index.html"} {small} {strong} {font face="Bradley Hand ITC"} Scholarships {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="sdp.htm"} {small} {strong} {font face="Bradley Hand ITC"} Senior Engineering Design Projects {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="stuorg.htm"} {small} {strong} {font face="Bradley Hand ITC"} Student Org. {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="21"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="http://www.wmich.edu/sunseeker/"} {small} {strong} {font face="Bradley Hand ITC"} Sunseeker {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="23"} {!--mstheme--} {font face="times new roman, times new roman, times"} {a href="latest.htm"} {small} {strong} {font face="Bradley Hand ITC"} The Latest {/font} {/strong} {/small} {/a} {!--mstheme--} {/font} {/td}
{/tr}
{tr}
{td align="center" width="469" height="1"} {!--mstheme--} {font face="times new roman, times new roman, times"} {!--mstheme--} {/font} {/td}
{/tr}
{/table} {!--mstheme--} {font face="times new roman, times new roman, times"}
{/div} {div align="left"}

{!--mstheme--} {/font} {table border="0" cellpadding="0" cellspacing="0" width="140" height="1"}
{tr}
{td width="136" height="1"} {!--mstheme--} {font face="times new roman, times new roman, times"}    
                    
              {!--mstheme--} {/font} {/td}
{/tr}
{/table} {!--mstheme--} {font face="times new roman, times new roman, times"}
{/div}
{!--mstheme--} {/font} {/td} {td valign="top" width="24"} {/td} {!--msnavigation--} {td valign="top"} {!--mstheme--} {font face="times new roman, times new roman, times"}
{b}

{h3 ALIGN="CENTER"} {!--mstheme--} {font color="#3399FF"} Dean, College of Engineering and Applied Sciences {br}
Western Michigan University {!--mstheme--} {/font} {/h3}
{/b}

{p} {small}     Western Michigan University seeks applications and
nominations for the position of Dean of the College of Engineering and Applied Sciences. {/small} {/p}

{p} {small}     Western Michigan University is a Carnegie Doctoral I
university with 750 FTE tenure-track faculty and an enrollment of 26,500 students, 25% at
the graduate level. It is advancing to Carnegie Research II classification, and plans a
substantial investment in the College of Engineering and Applied Sciences, including a new
physical facility, to accomplish this goal. In addition to its Graduate College and Lee
Honors College, Western supports six degree-granting colleges: Arts and Sciences, Haworth
College of Business, Education, Engineering and Applied Sciences, and Health and Human
Services. These colleges offer 242 academic programs, including 25 at the doctoral and 60
at the master's levels. {/small} {/p}

{p} {small}     The Dean of Engineering and Applied Sciences, who will report
to the Provost, must possess the vision that will enable the college to reach its fullest
potential in teaching, research and the development of industry partnerships, and also the
passion to implement this vision. The dean must demonstrate significant leadership
experience, knowledge of current and future issues in engineering education, a record of
successful participation in sponsored research, an ability to interact effectively with
external constituents, strong communication skills, and possess academic credentials that
would qualify for an appointment as professor with tenure in one of the College's
departments. {/small} {/p}

{p} {small}     The individual selected will assume the academic and
administrative leadership of a dynamic and growing college offering 21 baccalaureate, ten
masters and three doctoral programs. The college's staff, which currently includes 77
full-time faculty, 25 funded staff positions, and dozens of contract staff and graduate
student assistants, teaches more than 3000 students and generated $16.5M in contract
research in 1997-98. The College has a strong commitment to research and education, which
spans a broad spectrum of engineering and engineering technologies. In addition, it
pursues major research and training efforts in printing and in the aviation sciences. The
College offers courses and programs primarily on Western Michigan University's main campus
in Kalamazoo, but also serves as a major resource for off-campus instruction and economic
growth at five sites across West Michigan. {/small} {/p}

{p} {small}     WMU's main campus is located just off Interstate 94 and U.S.
Highway 131 in the southwest Michigan city of Kalamazoo, which is less than three hours by
car from both Detroit and Chicago. With a population of 220,000, Kalamazoo County is
served by ample air, train and bus transportation and offers an appealing environment for
study, employment, culture, entertainment and all-season recreation. {/small} {/p}

{p} {small}     For additional information about WMU and the College of
Engineering and Applied Sciences, refer to our Website: {i} http://www.wmich.edu/ {/i} {/small} {/p}

{p} {small} Applications and nominations should be directed to {/small} {/p}

{p ALIGN="CENTER"} {small} Dean James W. Schmotter, Chair, {/small} {/p}

{p ALIGN="CENTER"} {small} College of Engineering and Applied Sciences Dean Search Committee {/small} {/p}

{p ALIGN="CENTER"} {small} Haworth College of Business {/small} {/p}

{p ALIGN="CENTER"} {small} Western Michigan University {/small} {/p}

{p ALIGN="CENTER"} {small} 1201 Oliver Street {/small} {/p}

{p ALIGN="CENTER"} {small} Kalamazoo, MI 49008 {/small} {/p}

{p ALIGN="CENTER"} {small}   {/small} {/p}

{p} {small} For fullest consideration, applications and nominations should be received
before January 31, 1999. {/small} {/p}
{b}

{p ALIGN="CENTER"} WESTERN MICHIGAN UNIVERSITY IS AN {/p}

{p ALIGN="CENTER"} EQUAL OPPORTUNITY/AFFIRMATIVE ACTION EMPLOYER {/b} {/p}

{p ALIGN="CENTER"}   {/p}

{p}   {/p}

{p}  
{!--mstheme--} {/font} {!--msnavigation--} {/td} {/tr} {!--msnavigation--} {/table} {!--msnavigation--} {table border="0" cellpadding="0" cellspacing="0" width="100%"} {tr} {td} {!--mstheme--} {font face="times new roman, times new roman, times"}
{div align="left"}

{!--mstheme--} {/font} {table border="0" cellpadding="0" cellspacing="0" width="800" height="1"}
{tr}
{td height="1" width="171"} {!--mstheme--} {font face="times new roman, times new roman, times"} {!--mstheme--} {/font} {/td}
{td height="1" width="588"} {!--mstheme--} {font face="times new roman, times new roman, times"} {p align="center"} {img src="_borders/wmubar.jpg" WIDTH="429" HEIGHT="21"} {/p}
{p align="center"} Comments? broken link(s)?  {a href="mailto:s8Jasnie-at-unix.cc.wmich.edu"} click here {/a} {!--mstheme--} {/font} {/td}
{/tr}
{/table} {!--mstheme--} {font face="times new roman, times new roman, times"}
{/div}
{!--mstheme--} {/font} {/td} {/tr} {!--msnavigation--} {/table} {/body}
{/html}

--------------C23593DE2542DC81AE91F209--






From: LI Kun :      k-li-at-imre.org.sg
Date: Wed, 9 Dec 1998 11:13:34 +0800
Subject: Wulff net

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Could somebody please tell me where Wulff net can be downloaded?

Thanks for your help in advance.

Kun Li

Kun Li, Ph. D
Institute of Materials Research and Engineering
BLK S7, Level 3
Office: BLK S13, #02-13d
National University of Singapore
Lower Kent Ridge Road
Singapore 119260

Tel: 65-874 8187(Office); 65-874 3253(TEM Lab); 65-874 2999(Surface Lab)
Fax: 65-872 0785; e-mail: k-li-at-imre.org.sg.






From: CMontana4-at-aol.com
Date: Wed, 9 Dec 1998 00:27:41 EST
Subject: Re: W etches

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Marisa!
Depending on how much you want to etch, there are two that I've used to de-
process semiconductors. Typically I'm shooting to stop on a TiN liner or Si -
I use incremental times with these etches and check with a SEM to determine
the depth of the etch (top-down and simple cleaved sections are great for this
study).

1). Mix a 1:1 solution of NH4OH to H2O2 - both should be reagent grade, and
the H202 about 35%. The tungsten will fizz while under attack - experiment
with 2-10 minutes depending on your thickness. Use within 2 hours of mixing
for best reliability - the H2O2 likes to decompose.
2). Use household bleach (sodium hypochlorite) at 60 degrees C for about the
same amount of time as etch #1. You can purchase a reagent grade for more
money, but this does the trick too!

I do a dip rinse in a beaker, then hit it with a stream of water, then blow
dry with N2. Ultrasonic cleaning is fine if your structure can stand it, but
is usually not required.

Lisa Montanaro
Consultant
Microscopy/Microscopy Education
ph: (703) 257-7157
fax: (703) 365-2427
e-mail: cmontana4-at-aol.com

In a message dated 12/8/98 11:18:05 PM Mid-Atlantic Standard Time,
mahmad-at-semiconductor.com writes:

{ { From: mahmad-at-semiconductor.com (Marisa Ahmad)
To: Microscopy-at-sparc5.microscopy.com ('MSA listserver')


Anyone out there know of a nice uniform tungsten etch, preferably that will
stop on TiN/Ti or SiO2???? Dry or wet would be fine...any suggestions will
be much much much appreciated.

Thanks in advance,

Marisa Ahmad : )
} }





From: Randy Tindall :      rtindell-at-NMSU.Edu
Date: Tue, 08 Dec 1998 23:42:10 -0700
Subject: Re: Salary survey-Enough Already!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Granted that the salary survey data may not be sufficient for statistical
purposes, granted that we all make more than we really need (I'm an old
Peace Corps person, so please don't even try to convince me that we're
poor), granted that there are questions concerning time of entry into the
field, years of service, etc., I think that this is an ideal forum for such
a discussion. We are, after all, microscopists and this is, after all, a
forum for people like us.

Salary inequities between men and women are not confined to our field.
There may always be disagreement about what the figures actually PROVE, but
I simply find it amazing that people continue to debate that salaries are
unaffected by gender. I'm old enough to have watched this debate unfold
for a couple of decades and the story remains the same---women protesting
that salaries are unequal and some men protesting (with their higher
salaries) that everything is fine and that the problem is with the data
collection. Please note that I said "some men". Please note also that I am
a white, American euro-male on the wrong side of the Political Correctness
spectrum. Please note finally that I have NEVER been paid or offered
anywhere near close to the average salary quoted for either men or women in
this field, even though I have just been offered a position at a
considerably higher salary than I make now, after 12 years in the field.
It is still considerably below the average posted for female microscopists.
Am I complaining? No.

The story I hear is that women are paid less than men in many fields. Some
protest that this is because of late entry by women into these fields. So,
are there any figures out there for people in their first, fifth, or tenth
year in the field, correlated against educational background?

Let's continue to talk about this. It's as relevant as any other
microscopy topic. If their truly is unfairness in pay scales, it should be
brought to light and fixed by all means possible. IMHO.

Regards,
Randy





Randy Tindall
Las Cruces, New Mexico







From: Emilio Pastore :      epastore-at-ux1.unipd.it
Date: Wed, 9 Dec 1998 10:10:40 +0100 (MET)
Subject: unsubscrive stop

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscrive






From: Emilio Pastore :      epastore-at-ux1.unipd.it
Date: Wed, 9 Dec 1998 10:17:40 +0100 (MET)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe






From: Susan Cooper (Anatomiese Patalogie) :      gnapsc-at-med.uovs.ac.za
Date: Wed, 9 Dec 1998 11:47:45 GMT+2
Subject: Need info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


TEM --- Fixation and processing of tick eggs.

Any suggestions on how to prepair tick eggs successfully for TEM ? Poor
penetration seems to be the main problem because the cuticle is very hard.
I would appreciate your expert advise very much.

Thanks

Susan Cooper
Mrs Susan Cooper
Electron Microscope Unit
Department of Anatomical Pathology
University of the Orange Free State
P.O.Box 339 (G5)
Bloemfontein 9300
South Africa
Fax:0027-51-4473222 Tel:0027-51-4053061
E-Mail: GNAPSC-at-MED.UOVS.AC.ZA





From: Rik Brydson :      MTLRMDB-at-LUCS-MAC.NOVELL.LEEDS.AC.UK
Date: Wed, 9 Dec 1998 11:06:31 GMT0BST
Subject: SuperSTEM position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Any people involved with TEM or STEM on the list(s) may be interested in this.

SuperSTEM project

Invitation to Interested Parties for Possible One Year Post

With current advances in lens aberration correctors occurring worldwide, the Electron
Microscopy and Analysis (EMAG) committee of the Institute of Physics (IOP) is currently
investigating the possibility of setting up an analytical TEM/STEM/ PEELS facility for the UK
and European microscopy community. The facility is envisaged to consist of at least two
machines capable of high resolution imaging and analysis at the sub-Angstrom level and
would provide a user service to external researchers in parallel with constant development in
instrumentation. From these discussions, the possibility of an initial one year post, most
probably located at Daresbury Laboratories in Cheshire, has arisen. This appointment would
enable the project to be throughly researched and prepared for funding submission. The
SuperSTEM sub-committee is keen to hear from interested parties and potential applicants for
this post. In the first instance, candidates are asked to contact Prof. L.M. Brown. (0044
(0)1223 337291, email: lmb12-at-cam.ac.uk)



_____________________________
Dr. Rik Brydson,
University Research Fellow,
Electron Optical Unit,
Department of Materials,
School of Process, Environmental and Materials Engineering
University of Leeds,
Leeds LS2 9JT,
U.K.

Tel: 44 + (0)113 233 2369
Fax: 44 + (0)113 242 2531
Web: http://www.leeds.ac.uk/materials

______________________________

**************************
Join email discussion list for all aspects of
Electron Microscopy and Analytical Spectroscopy (LEMAS)

send message "join lemas firstname lastname"
to mailbase-at-mailbase.ac.uk
**************************





From: Gerd Duscher :      duscher-at-hrem.mpi-stuttgart.mpg.de
Date: Wed, 09 Dec 1998 13:32:49 +0100
Subject: Re: Magnification calib in STEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hallo Frank,
what kind of STEM do you have?
We use Si lattice images for the magnification calibration at
the VG HB501 UX.
If you don't get this resolution,
you could use something with a large lattice spacing.
Gerd

--
*******************************************************************************
Dr. Gerd Duscher
MPI fuer Metallforschung Tel.: ++49 711 2095 313
Institut fuer Werkstoffwissenschaft Fax : ++49 711 2095 320
Seestr. 92 e-mail: duscher-at-hrem.mpi-stuttgart.mpg.de
D-70174 Stuttgart
*******************************************************************************





From: Colin Reid :      creid-at-tcd.ie
Date: Wednesday, December 09, 1998 12:09 PM
Subject: Re: Salary survey-Enough Already!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


---------------------- Forwarded by Pat Kingman/arl on 12/09/98 07:45 AM
---------------------------


Pat Kingman
12/09/98 08:02 AM
To: LI Kun {k-li-at-imre.org.sg}
cc:


Hey Randy don't disillusion me. I was just about to arrange a sex change &
apply for a green card !!

Colin


Colin Reid,
Electron Microscope Unit,
Trinity College Dublin,
Dublin 2,
Ireland.
Tel: 353-1-6081820
Fax: 353-1-6770438
email: creid-at-tcd.ie
-----Original Message-----
} From: Randy Tindall {rtindell-at-NMSU.Edu}
To: microscopy-at-sparc5.microscopy.com {microscopy-at-sparc5.microscopy.com}






From: Nestor J. Zaluzec :      zaluzec-at-Sparc5.Microscopy.Com
Date: Wed, 9 Dec 1998 08:31:17 -0500
Subject: Administrivia: Salaries...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues...

Discussion on salaries as it relates to experise in microscopy
is in my view within the scope of this forum. It is abit off the
main line of microscopy/microanalysis per se, but I think of
sufficient interest to the community that if someone has a
constructive comment or information it should be posted.

However, I think we have had more than enough of the side
jokes, puns, etc. for now.

My guess is that most people have had their say already.

Your Friendly Neighborhood SysOp

Nestor







From: Doug Cromey :      doug-cromey-at-ns.arizona.edu
Date: Wed, 09 Dec 1998 07:42:14
Subject: ? mounting renal tubules for immuno

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have an interesting technical problem that I'm hoping the folks on this
list can help me figure out.

I have a colleague who is trying to embed dissected renal tubules (20
micron diameter x 500 micron length) for immunofluorescence. He would like
to be able to "tack down" individual tubules to some kind of substrate, so
that he can keep track of which end he's working from, and would like to
cross section them (1-4 micron sections) from end-to-end (vs longitudinal
sx). We've tried adhering tubules to Thermanox coverslips and making
cryotome sections, but the sections of plastic curled up like a scroll.
We've tried Aclar film, and plastic embedding using LR Gold and had the
same problem. Is there someone out there who has a suggestion? We have 2
concerns, 1) we don't want to heat the specimen above physiologic temps and
2) we want to keep track of orientation so as to have cross sections of the
tubules - adherence to some sort of substrate is one possible solution -
any alternatives?

Thanks for your help.
Doug
....................................................................
: Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy :
: Research Specialist, Principal University of Arizona :
: (office: AHSC 4212A) P.O. Box 245044 :
: (voice: 520-626-2824) Tucson, AZ 85724-5044 USA :
: (FAX: 520-626-2097) (email: doug-cromey-at-ns.arizona.edu):
:...................................................................:
http://www.pharmacy.arizona.edu/exp_path.html
Home of: "Microscopy and Imaging Resources on the WWW"






From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Wed, 09 Dec 1998 10:42:33 -0800
Subject: Re: Need info/tick egg fixation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Susan Cooper (Anatomiese Patalogie) wrote:
}
} TEM --- Fixation and processing of tick eggs.
} Any suggestions on how to prepair tick eggs successfully for TEM ? Poor penetration seems to be the main problem because the cuticle is very hard.
} I would appreciate your expert advise very much.
}
Phase-partition fixation would be worth a try, often useful for
invertebrate eggs. See Zalokar and Erk, 1977, Stain Technology 52:89.
"Phase partition fixation and staining of Drosophila eggs."

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (732)-235-4583; fax -4029 e-mail: mcauliff-at-umdnj.edu
***************************************************************





From: Giles, John E Jr (FL51) :      jegiles-at-space.honeywell.com
Date: Wed, 9 Dec 1998 10:34:19 -0500
Subject: Need Ideas for Analysis of Conductive Epoxy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


All,

I have been asked about analyzing a silver filled (70%) conductive epoxy to
look for evidence of a lubricant or other material on the surface of the
silver flakes in the epoxy. The epoxy is exhibiting variations in
conductivity and one of the hypotheses is that there is a lubricant or other
material present that may be influencing conductivity. There is much more
to the story, but I won't go into that in order to keep the message short.

We have done quite a bit of ESCA and SEM work on the bulk material, but
believe we may need to look at the material in a different manner. One of
the suggestions has been to microtome the sample and do TEM on it. The
thought is we may be able to see the lubricant layer surrounding the silver
flakes at high magnifications. I am only slightly familiar with TEM (I'm a
SEM and Surface Analysis person) and thought I would approach the group and
ask about the feasibility of this method or any other suggestions would be
most welcome. We have the material available in the cured and uncured
states.

Thanks in advance for any suggestions,

John Giles
Principal Materials Engineer
Honeywell Space Systems







From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Wed, 09 Dec 98 10:48:57 -0500
Subject: Wulff projections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Kun Li wrote:
===============================================
Could somebody please tell me where Wulff net can be downloaded?
================================================
I don't know of any place where they could be "downloaded" but they can be
purchased from our firm, SPI Supplies. I presume they can be purchased from
others, possibly EBS. You can find out details of these and other crystal
plane projections on our website or go directly to URL
http://www.2spi.com/outlet/project1.html

Chuck

===================================================
Charles A. Garber, Ph. D. Ph: 1-(610)-436-5400
President 1-(800)-2424-SPI
SPI SUPPLIES FAX: 1-(610)-436-5755
PO BOX 656 e-mail: cgarber-at-2spi.com
West Chester, PA 19381-0656 USA Cust. Service: spi2spi-at-2spi.com


Look for us!
############################
WWW: www.2spi.com
############################
==================================================





From: rick-at-pgt.com (Rick Mott)
Date: Wed, 9 Dec 98 11:13:51 EST
Subject: creation of Lemas list not considered harmful

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



A couple of negative comments have been posted about the creation of
the Lemas list, but no similar fuss was raised when the microprobe
list was created which also has considerable overlap with the MSA
listserver.

There's no harm in also having lists with a somewhat narrower focus.
I'll continue to subscribe to all three and suffer some duplication
of messages, but I can certainly understand someone with a materials
science interest not wanting to wade though all the biological sample
prep material which often dominates the microscopy listserver.

There's not much difference between sorting postings by subject in
one listserver and subscribing to multiple listservers, except the
latter at least offers the option of not receiving mail which isn't
of interest. For folks who pay for their mail messages, that might
be significant.

Rick Mott
rick-at-pgt.com





From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Wed, 09 Dec 1998 11:30:58 -0500
Subject: Re: Need info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We had a similar problem with wasp eggs and posted the question to the
list. The discussion is archived at the following url at the "Tips &
Tricks" site.

http://www.biotech.ufl.edu/icbr/emcl/db/insect.html

other useful links include;

http://www.biotech.ufl.edu/icbr/emcl/db/roots.html

Good luck





At 11:47 AM 12/09/1998 GMT+2, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America




} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {
GO GATORS
Scott D. Whittaker 218 Carr Hall
EM Technician Gainesville, FL 32610
University Of Florida ph 352-392-1184
ICBR EM Core Lab fax 352-846-0251
sdw-at-biotech.ufl.edu http://www.biotech.ufl.edu/~emcl/
The home of " Tips & Tricks "










From: Caroline Schooley :      schooley-at-mcn.org
Date: Wed, 9 Dec 1998 09:08:04 -0800 (PST)
Subject: Project MICRO

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi folks -

I have news about two useful supplements for MSA's middle school (grades
4-8) manual,"Microscopic Explorations":

Warren Hatch is a Portland, Oregon schoolteacher who makes very good
microscopy videos for children. They're cheap ($15-30 postpaid, depending
on quantity, format, and destination); he makes them as a labor of love.
MICRO asked him to make a "custom" tape to match the exercises in
"Microscopic Explorations". It's now ready, and it's as good as his other
tapes. It's 33 minutes long & it's available in both VHS and PAL. Contact
him directly at {whatch-at-hevanet.com} for ordering information.

If you are already using "Microscopic Explorations" you know that it
contains copyable student instructions and worksheets. The Lawrence Hall of
Science is now completing a full set of these in Spanish. If you could use
them, or know someone who could, please let me know.

Caroline Schooley
Educational Outreach Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.MSA.microscopy.com/ProjectMICRO/Books.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html







From: Heinz Fehrenbach :      hefeh-at-Rcs1.urz.tu-dresden.de
Date: Wed, 9 Dec 1998 18:14:51 +0100
Subject: Re: Need info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear Susan,

some years ago, I did a lot of work on TEM of lepidopteran
eggs, which are likewise enclosed in a +/- thick shell.
Best fixation was achieved, although not in every single
egg, but in most of them, by pricking the eggshell with
a tiny glass needle that you make from a glass capillary.
Be sure that the needle has a sharp edge by breaking it
with a Dumont forceps (size 7).

As a primary fixative, we used a mixture of glutaraldehyde
and osmium tetroxide in cacodylate buffer. Fixation was
done on ice for 30-60 minutes. After several rinses, post-
osmication was performed followed by dehydration and Epon
embedment - see H. Fehrenbach (1995) Zool. Anz. 234:19-41 and
D. Zissler, K. Sander (1973) W. Roux' Arch. Entw.mech. 172:175-86.

Hope that helps.

Good luck with your work,

Heinz



-----------------------------------------------------------------
Dr. Heinz Fehrenbach
Institute of Pathology
University Clinics "Carl Gustav Carus"
Technical University of Dresden

Fetscherstr. 74 Phone: ++49-351-458-5277
D-01307 Dresden Fax: ++49-351-458-4328
Germany e-mail: hefeh-at-rcs.urz.tu-dresden.de
-----------------------------------------------------------------







From: Dr. Kalpana S Katti :      kkatti-at-prairie.nodak.edu
Date: Wed, 9 Dec 1998 11:30:06 -0600 (CST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Unsubscribe







From: hank p adams :      hpadams-at-bcm.tmc.edu
Date: Wed, 9 Dec 1998 12:29:02 -0000
Subject: Salaries, again

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I originally asked about 10 months ago if anyone was aware of a recent =
salary survey for microscopists similar to the one that was conducted by =
MSA back in 1984. I had been asked by the Vice Pres of Research and the =
Personnel Director at the institution I was employed at, previously, =
what were the salaries paid microscopists in reply to my asking for an =
adjustment based on my responsibilities and years of experience. ( I =
found that the earlier survey by MSA was helpful in a salary adjustment =
back in 1988, but now was invalid due to its age.) I believe these =
surveys serve a useful purpose as it exposes in equities (as has been =
brought out here) and also give us a benchmark when hiring new people, =
and are helpful to personnel directors in determining wage scales. =
Furthermore there is a significant difference in cost of living through =
out the US, which surveys address as well as specialties, etc. =20
I know some fields involving microscopy are experiencing a decline in =
demand but the demand in some other areas for microscopists trained in =
EM and in the new light microscopies with experience in networking and =
image processing has grown. These labs are now looking for people with =
more capabilities/experience than when I started working in this field =
in the mid seventies and administers/personnel directors expect us to =
hire them at salaries that do not reflect their skills. Any new, valid =
data would be helpful, unfortunately, not enough people responded to =
give us current and valid data.
Hank Adams
Cell Biology
Integrated Microscopy Core
Baylor College of Medicine
One Baylor Plaza
Houston, Tx 77030






From: CORLB-at-polaroid.com (R-Brooks Corl)
Date: Wed, 09 Dec 1998 10:56:50 -0500
Subject: Looking for Time Lapse Software Source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Several months ago I was told about a software developer who has some
relatively simple "time-lapse" imaging acquisition programs for
digital cameras. I have since lost track of the name and source.
Does anyone know who this is?





From: HILDEGARD CROWLEY :      hcrowley-at-du.edu
Date: Wed, 09 Dec 1998 11:42:07 -0700 (MST)
Subject: Salaries,M,Fe,US Dep Labor Stat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In the discussion of salaries for Microscopists we must include an
extremely informative bit of information put out by the US Dep of Labor
in 1983. That is a long time ago now, but nevertheless, it needs to be
known. At the time of the statistics, Fe salaries were 70 cents or so to
M sal of 1.00. This was widely published ( and much screamed about) in
every newspaper, etc. The US Dep of Labor from the census divided men and
women into 4 groups and averaged their salaries. The four groups: Married
Men, Unmarried Men, Married Women, Unmarried Women. The results were as
follows:
Married Men had the highest salaries. Unmarried Women had slightly,
slightly, slightly less salary. It was approximately 6% less. Then there
was a considerable drop for unmarried men. There followed a HUGE drop for
Married women. Married women made about half of what married men did.
The commentary at that time was (and some of that is still true) that 1)
Married women consider their income frequently as a supplement to the
family income and work part time while raising the children and running
the house. Many women still follow this pattern. Also many women still
take off several years while children are young, or follow less ambitious
career paths for several years. We do need to consider this in the
overall picture.
In 25 years of working professionally I have watched men VERY closely
(fun). To see what they do and how they do it. Generally, not always,
men are MUCH more forceful when asking for raises, or making moves to
other towns for more power and more money than women. Men will write
letters, threaten the department chair with leaving, endlessly scout the
scene for better jobs, etc. Women are just now learning this type of
behavior. Women who demand do a lot better than those who don't.
It will change. It is changing now. There will be different ways of
living in the next two generations. It will be exciting and interesting.
Bye,
hildy






From: Bob Lawrence (a402aa) :      a402aa-at-email.sps.mot.com
Date: Wed, 09 Dec 1998 13:11:47 -0700
Subject: Re: creation of Lemas list not considered harmful

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Rick,

I was going to say something like this, but I couldn't have said it
any better so I will just say " Thank You!!" I am primarily a materials
science man myself, with secondary interests and occasional assignments
in the bio area. I find I never have enough good information and I can
deal with overlap.

--
Respectfully,
Bob ( Robert G. ) Lawrence
Failure Analyst
Motorola Phoenix Corporate Research Lab
2100 E. Elliot Rd.
MD 508
Tempe, AZ 85284-1806
Phone: 602-413-5848
Fax: 602-413-5934
Pager: 1-800-759-7243
PIN 834-2458







From: hank p adams :      hpadams-at-bcm.tmc.edu
Date: Wed, 9 Dec 1998 14:21:03 -0000
Subject: Salaries, again.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I originally asked about 10 months ago if anyone was aware of a recent =
salary survey for microscopists similar to the one that was conducted by =
MSA back in 1984. I had been asked by the Vice Pres of Research and the =
Personnel Director at the institution I was employed at, previously, =
what were the salaries paid microscopists in reply to asking for an =
adjustment based on my responsibilities and years of experience. ( I =
found that the earlier survey by MSA was helpful in a salary adjustment =
back in 1988, but now was invalid due to its age.) I believe these =
surveys serve a useful purpose as it exposes in- equities (as has been =
brought out here) and also give us a benchmark when hiring new people, =
and are helpful to personnel directors in determining wage scales. =
Furthermore there is a significant difference in cost of living through =
out the US, which surveys address as well as specialties, etc. =20
I know some fields involving microscopy are experiencing a decline in =
demand but the demand in some other areas for microscopists trained in =
EM and in the new light microscopies with experience in networking and =
image processing has grown. These labs are now looking for people with =
more capabilities/experience than when I started working in this field =
in the mid seventies and administers/personnel directors expect us to =
hire them at salaries that do not reflect their skills. Any new, valid =
data would be helpful, unfortunately, not enough people responded to =
give us that. Thanks, to Don Grimes though for making an attempt. Maybe, =
M&M should conduct one, since the only data available is from 1984.


Hank Adams
Cell Biology
Integrated Microscopy Core
Baylor College of Medicine
One Baylor Plaza
Houston, Tx 77030






From: Brendalyn Bradley :      bradley_kerr-at-ncsu.edu
Date: Wed, 9 Dec 1998 15:15:37 -0500 (EST)
Subject: LM-GMS staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for a reliable procedure for staining tissue embedded in glycol
methacrylate(JB-4 resin) with silver (GMS).


Brendalyn Bradley-Kerr
LAELOM
College of Veterinary Medicine
North Carolina State University
Raleigh, NC 27606






From: Barbara Foster :      mme-at-map.com
Date: Wed, 09 Dec 1998 17:18:36 -0500
Subject: Re: Need info

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,


I am going to make the assumption that since you are resorting to TEM you
need to resolve extremely detail within the egg. Just in case that is
not the situation, I once did a neat set of confocal images of a moth
egg, both inside and out, using confocal. We did the outside using
reflected light and the inside using autofluorescence. No fixation was
necessary and the results were amazingly detailed. Just a thought.


Best regards,

Barbara Foster

Consortium President

{color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
..Educating microscopists for greater productivity.


{/color} 125 Paridon Street Suite 102 Springfield, MA 01118

PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

Visit our web site { {http://www.MME-Microscopy.com/education}

******************************************************

{bigger} {bigger} MME {/bigger} {/bigger} is America's first national
consortium dedicated to

customized on-site training in all areas of

microscopy, sample preparation, and image analysis {bigger} .

{/bigger}




At 06:14 PM 12/9/98 +0100, Heinz Fehrenbach wrote:

} ------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

} -----------------------------------------------------------------------.

}

}

}

} Dear Susan,

}

} some years ago, I did a lot of work on TEM of lepidopteran

} eggs, which are likewise enclosed in a +/- thick shell.

} Best fixation was achieved, although not in every single

} egg, but in most of them, by pricking the eggshell with

} a tiny glass needle that you make from a glass capillary.

} Be sure that the needle has a sharp edge by breaking it

} with a Dumont forceps (size 7).

}

} As a primary fixative, we used a mixture of glutaraldehyde

} and osmium tetroxide in cacodylate buffer. Fixation was

} done on ice for 30-60 minutes. After several rinses, post-

} osmication was performed followed by dehydration and Epon

} embedment - see H. Fehrenbach (1995) Zool. Anz. 234:19-41 and

} D. Zissler, K. Sander (1973) W. Roux' Arch. Entw.mech. 172:175-86.

}

} Hope that helps.

}

} Good luck with your work,

}

} Heinz

}

}

}

} -----------------------------------------------------------------

} Dr. Heinz Fehrenbach

} Institute of Pathology

} University Clinics "Carl Gustav Carus"

} Technical University of Dresden

}

} Fetscherstr. 74 Phone: ++49-351-458-5277

} D-01307 Dresden Fax: ++49-351-458-4328

} Germany e-mail: hefeh-at-rcs.urz.tu-dresden.de

} -----------------------------------------------------------------

}

}

}

}

}







From: Tim P. LaFave Jr. :      lafatim-at-charlie.cns.iit.edu
Date: Wed, 9 Dec 1998 19:35:41 -0600 (CST)
Subject: Re: creation of Lemas list not considered harmful

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


After reading through some of these listserver segregation msgs I have
thought of an idea, which I'm sure someone has constructed before, or
perhaps not (even possible). However, here it is:

The listserver should allow users to subscribe to a handful of segregated
topical specializations. YOu may choose any number of them. However, when
posting the listserver should offer the possibility of posting to all
groups, or just a select number of them. Each group will automatically
serve onto itself(a person responding to materials will only be posted to
materials automagically), however, the listserver should be set up such
that
a user can post to all groups (to avoid duplication, etc. to people have
subscr.d to more than one). The listserver may be set up in such a way
that if a user wishes to post, he/she would have to list the groups' names
in the subject line, or perhaps the first line of the message, according
to which groups the message is to be posted. Perhaps a simple code of
"abcd" or "xyzt" could be listed in the subject line, which will tell the
listserver to post to all groups, or any variation therein for separate
groups. This will save resources and resolve the issue of segregating the
microscopy listserver...we could even begin to bring other slightly
-removed topics into the listserver with absolutely no one complaining
(unless the listserver goes nuts).

The net effect is that everyone can subsr. to all groups (or just the ones
they want). When a person wants to post with a general request, e.g.
corporate, academic, or other general umbrella topics, the msg can be
posted to any number in the set of {xyzt...}. Effectivley no one loses
content, no one receives duplication, and everyone can post to all groups.
(HENCE EVERYONE IS SATISFIED AND RESOURCES ARE SPARED)

..but I'm not sure how a listserver can be set up to do this...I'm sure
it has been done somewhere. ...and shouldn't be a complex task to
implement.


[ignore the name below, as I'm not quite willing to write the listserver
code]
__ _-==-=_,-.
/--`' \_-at---at-.-- {
Tim (TJ) LaFave Jr. `--'\ \ {___/.
Department of Physics \ \\ " /
University of North Carolina, Charlotte } =\\_/` {
Charlotte, NC 28223 ____ /= | \_|/
_' `\ _/=== \___/
(704)547-3392 [x4] `___/ //\./=/~\====\
(704)509-6622 [Hm] \ // / | ===:
http://www.iit.edu/~lafatim | ._/_,__|_ ==: __
\/ \\ \\`--| / \\
---------- +*+ ---------- | _ \\: /==:-\
`.__' `-____/ |--|==:
Such that the future be theirs \ \ ===\ :==:`-'
to shape and direct. _} \ ===\ /==/
-----------------------------------------------------------------

On Wed, 9 Dec 1998, Rick Mott wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
}
} A couple of negative comments have been posted about the creation of
} the Lemas list, but no similar fuss was raised when the microprobe
} list was created which also has considerable overlap with the MSA
} listserver.
}
} There's no harm in also having lists with a somewhat narrower focus.
} I'll continue to subscribe to all three and suffer some duplication
} of messages, but I can certainly understand someone with a materials
} science interest not wanting to wade though all the biological sample
} prep material which often dominates the microscopy listserver.
}
} There's not much difference between sorting postings by subject in
} one listserver and subscribing to multiple listservers, except the
} latter at least offers the option of not receiving mail which isn't
} of interest. For folks who pay for their mail messages, that might
} be significant.
}
} Rick Mott
} rick-at-pgt.com
}






From: Dr. Mark W. Lund :      lundm-at-physc3.byu.edu
Date: Wed, 09 Dec 1998 22:01:32 MST/MDT
Subject: Lemas List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My observations after being on 10 or so lists over the years
is that you need a critical mass of active subscribers to
make the thing self sustaining. Unless you get a steady
stream of posts the list is quickly forgotten. This list
has it. The microbeam analysis list doesn't. Any more
fragmenation of the list makes both parts less able to
survive.

My vote--the area is already being covered my a working list,
let's not mess with it.

Best regards
mark

Mark W. Lund, PhD
VP Engineering } } Soft X-ray Web page http://www.moxtek.com { {
MOXTEK, Inc.
452 West 1260 North
Orem UT 84057 801-225-0930 FAX 801-221-1121
lundm-at-xray.byu.edu

"This is a YOUNG business...How can I tell you what
YOUR job is when I don't know what MINE is?" --Pogo





From: Ellis, Sarah :      s.ellis-at-pmci.unimelb.edu.au
Date: Thu, 10 Dec 1998 16:18:32 +1100
Subject: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This message is in MIME format. Since your mail reader does not understand
this format, some or all of this message may not be legible.

------_=_NextPart_000_01BE23FC.8C865A50
Content-Type: text/plain;
charset="iso-8859-1"

Hello all again,


It seems that cost cutting and space saving have finally started attacking
us! Our darkroom, which I admit is rather large, is wanted by another more
influential person. It has been suggested that we attach a digital camera
to our Hitachi H600 TEM to take over from the film camera we currently use.
I have suggested that digital images are not yet able to produce images as
good as our film can and that the digital image can not be increased to a
much larger size without appreciable loss of resolution. Can any one please
help us find a solution as this same influential person also has their beady
little eyes on our laboratory!
We need to know
1. How easy is it to digitise the TEM without losing the ability to
still turn to film.
2. What sort of camera/system should we be looking at?
3. What sort of software is needed?
4. How much do these things cost?

Sarah Ellis

Trescowthick Research Centre
Peter MacCallum Cancer Institute
Locked Bag #1 A'Beckett Street
Melbourne 8006 Victoria
Australia

Phone +61-3-9656 1244
Fax +61-3-9656 1411
Email s.ellis-at-pmci.unimelb.edu.au


------_=_NextPart_000_01BE23FC.8C865A50
Content-Type: application/ms-tnef
Content-Transfer-Encoding: base64
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------_=_NextPart_000_01BE23FC.8C865A50--





From: Tom Murray :      tm8a-at-Virginia.EDU
Date: Thu, 10 Dec 1998 08:28:34 -0500
Subject: Re: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have mixed feelings about another list server but if they evolve in their
own separate ways there may be enough of a committed user base to support
both. In any event at least it has been discussed in depth here.

My only suggestions are:
1. that, as many have pointed out recently, this is an international list
so could people always put at least a basic address and there country of
origin on their mail? There are still times when someone is selling kit or
wanting local advice and I waste time checking their e-mail addresses etc in
the futile hope of guessing where they are.
2. If people do post to more than one list can they do it at the same time
so we can see it in the 'address to:' or 'copies to:' headers and then know
whether to delete duplicates?

thanks

Malcolm Haswell
Electron Microscopy
School of Health Sciences
Fleming Building
University of Sunderland
SUNDERLAND SR1 3SD
Tyne and Wear
UK

Tel (0191) 515 2872
e-mail: malcolm.haswell-at-sunderland.ac.uk

----------
} From: Rick Mott
To: microscopy


} Hello all again,
}
}
} It seems that cost cutting and space saving have finally started attacking
} us! Our darkroom, which I admit is rather large, is wanted by another more
} influential person. It has been suggested that we attach a digital camera
} to our Hitachi H600 TEM to take over from the film camera we currently use.
} I have suggested that digital images are not yet able to produce images as
} good as our film can and that the digital image can not be increased to a
} much larger size without appreciable loss of resolution. Can any one please
} help us find a solution as this same influential person also has their beady
} little eyes on our laboratory!
} We need to know
} 1. How easy is it to digitise the TEM without losing the ability to
} still turn to film.
} 2. What sort of camera/system should we be looking at?
} 3. What sort of software is needed?
} 4. How much do these things cost?
}

One possibility is to use photoplates. I don't have any personal
experience with these, but my understanding is they are basically digital
film. The microscope does not need to be modified. Photoplates are used
in the same fashion as film. I understand they have a better dynamic range
than a CCD. They are also much higher resolution than a CCD. The down
side is cost. A reader costs in the $15,000 USD range I believe, and the
plates are close to $100 USD each. Also each image is 25 - 30 megabytes so
disk storage and backup for the images can add up in a hurry.

All of the above is from memory of a conversation with someone who had this
setup. The numbers may be misremembered. Hopefully someone who actually
has bought and used photoplates will respond.

Tom

Thomas Mullarkey Murray email:tm8a-at-virginia.edu
Thornton Hall - MSE phone:(804)982-5659
University of Virginia Fax: (804)982-5660
Charlottesville, VA 22903







From: Miller.David :      miller-at-niehs.nih.gov
Date: Thu, 10 Dec 1998 08:39:48 -0500
Subject: POSTDOCTORAL POSITION AVAILABLE

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


POSTDOCTORAL POSITION - CELL PHYSIOLOGY
I have an opening for a Postdoctoral Fellow in my laboratory . The research
is concerned with the cellular and molecular biology of drug transport
across the renal tubule and the blood-brain barrier. We use isolated renal
proximal tubules, renal cells in culture, isolated brain capillaries,
fluorescent substrates and confocal microscopy to follow transport across
cells. Our goals are to 1) characterize the fundamental membrane-based and
intracellular processes involved, 2) determine how they are controlled by
hormones and xenobioitcs, and 3) identify the signal transduction pathways
involved. Candidates should have a Ph.D. or M.D. degree, less than 5 years
of postdoctoral experience and a background in cellular physiology, membrane
transport or renal physiology. This is a non-tenure track position (NIH IRTA
Fellow).


_____________________________
Dr. David S. Miller
Laboratory of Pharmacology & Chemistry
NIH/NIEHS
Research Triangle Park, NC 27709

miller-at-niehs.nih.gov
919 541 3235





From: malcolm.haswell-at-sunderland.ac.uk (HASWELL Malcolm)
Date: 10 December 1998 16:18
Subject: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Sarah

I use a Hitachi H7000 and, although I don't have your urgent requirement, I
have looked at the convenience issue of digital photography on a TEM.

My understanding is that you would probably need to insert some sort of
fluorescent screen and high resolution digital camera at either the 35mm
camera port or STEM detector port position of the microscope. At present
this would either involve Hitachi or a specialist company that you could
trust. By the time that you have paid for all the specialist kit the numbers
I have been quoted don't come much below 30,000 UK pounds. It's probably
cheaper to get the system integrated into a new microscope if and when you
upgrade or have some spare petty cash you're not using.

However - have you considered a simpler and much cheaper option. Develop TEM
film in a simple darkroom (in an emergency any dark ventilated area will
do). Then scan negatives with the highest resolution negative/transparency
scanner you can get. I would guess that you could pay anything from 2,000 to
10,000 UK pounds depending on what you wanted and this would include
scanner, fast computer , storage, software and printer. The added advantage
is that from a typical negative and high resolution scanner it should be
possible if necessary to scan at much better resolutions than with the best
digital camera. I would guess that you would be able to print at up to at
least 5x the original negative magnification if you used the right scanner.

This is my master plan because you still have the archival and universal
properties of film but most of the convenience of digital plus you could at
some later date do a full up-grade. I'm still awaiting money to do this but
for the outlay the benefits greatly outweigh the costs.

Good luck and let me know how you get on.


Malcolm Haswell
Electron Microscopy
School of Health Sciences
Fleming Building
University of Sunderland
SUNDERLAND SR1 3SD
Tyne and Wear
UK

Tel (0191) 515 2872
e-mail: malcolm.haswell-at-sunderland.ac.uk
----------
} From: Ellis, Sarah
To: Microscopy Listserver

Hello all again,


It seems that cost cutting and space saving have finally started attacking
us! Our darkroom, which I admit is rather large, is wanted by another more
influential person. It has been suggested that we attach a digital camera
to our Hitachi H600 TEM to take over from the film camera we currently use.
I have suggested that digital images are not yet able to produce images as
good as our film can and that the digital image can not be increased to a
much larger size without appreciable loss of resolution. Can any one please
help us find a solution as this same influential person also has their beady
little eyes on our laboratory!
We need to know
1. How easy is it to digitise the TEM without losing the ability to still
turn to film.
2. What sort of camera/system should we be looking at?
3. What sort of software is needed?
4. How much do these things cost?

Sarah Ellis

Trescowthick Research Centre
Peter MacCallum Cancer Institute
Locked Bag #1 A'Beckett Street
Melbourne 8006 Victoria
Australia

Phone +61-3-9656 1244
Fax +61-3-9656 1411
Email s.ellis-at-pmci.unimelb.edu.au






From: Katherine.S.Connolly-at-dartmouth.edu (Katherine S. Connolly)
Date: 10 Dec 98 08:56:15 EST
Subject: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by mailhub.dartmouth.edu (8.8.8+DND/8.8.8) with SMTP id IAA30481
for {MICROSCOPY-at-MSA.Microscopy.Com} ; Thu, 10 Dec 1998 08:56:16 -0500 (EST)
Message-id: {12687646-at-vixen.Dartmouth.EDU}

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


We have been digitizing our images since 1995. Our solution to the massive
resolution loss in our biological digital images (301K) is to take films as
well (fewer) which serve as archives. My experience is you need 3 digitized
images at different magnifications to get out the same information contained in
one film. Films may be scanned as 3-4MB images and printed on a good 1200 dpi
dye-sub printer or printed the old-fashioned way.

Kate Connolly





From: Martin J. Roe :      mi596-at-mluri.sari.ac.uk
Date: Thu, 10 Dec 1998 15:26:10 +0000
Subject: Re: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by highgate.mluri.sari.ac.uk (8.8.7/8.8.7) with SMTP id OAA20601;
Thu, 10 Dec 1998 14:19:55 GMT
Message-Id: {199812101419.OAA20601-at-highgate.mluri.sari.ac.uk}
Comments: Authenticated sender is {mi596-at-highgate}


Hello Sarah,
You've asked an interesting question (at least to me that is).The
organisation I work for may be buying an old TEM in the near future
which could be upgraded to digital. So I'd be interested in hearing
all the responses. For example, do digital cameras really offer a
chance for us to get out and even stay out of the dark room?
(apologies if this question sounds simple). I know that one
would need a slow scan CCD rather than a TV rate digital
camera for suitable quality images.
I'd be interested in all your thoughts on the subject.
Regards

Martin Roe
MLURI
Craigiebuckler
Aberdeen
Scotland
U.K.

} It seems that cost cutting and space saving have finally started attacking
} us! Our darkroom, which I admit is rather large, is wanted by another more
} influential person. It has been suggested that we attach a digital camera
} to our Hitachi H600 TEM to take over from the film camera we currently use.
} I have suggested that digital images are not yet able to produce images as
} good as our film can and that the digital image can not be increased to a
} much larger size without appreciable loss of resolution. Can any one please
} help us find a solution as this same influential person also has their beady
} little eyes on our laboratory!
} We need to know
} 1. How easy is it to digitise the TEM without losing the ability to
} still turn to film.
} 2. What sort of camera/system should we be looking at?
} 3. What sort of software is needed?
} 4. How much do these things cost?
}
} Sarah Ellis
}
} Trescowthick Research Centre
} Peter MacCallum Cancer Institute
} Locked Bag #1 A'Beckett Street
} Melbourne 8006 Victoria
} Australia
}
} Phone +61-3-9656 1244
} Fax +61-3-9656 1411
} Email s.ellis-at-pmci.unimelb.edu.au
}
}
}





From: Robert Underwood :      underwoo-at-u.washington.edu
Date: Thu, 10 Dec 1998 06:55:04 -0800 (PST)
Subject: Re: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We only use a very small darkroom to develop the film. Basiclly a closet
next to the scope. Then we scan the negs at either 600 dpi or 1200dpi,
stor the files on CDs and print on good printers. We have gotten addicted
to
the enhanced information we get by viewing the images on the monitor, with
image enhancement and analysis tools at the click of a mouse. For us it
seems to be the best solution for now. We want to bypass film eventually
but the cost and drawbacks seem too great for now.

Bob
Derm Imaging Center
U of W

On Thu, 10 Dec 1998, Ellis, Sarah wrote:

} Hello all again,
}
}
} It seems that cost cutting and space saving have finally started attacking
} us! Our darkroom, which I admit is rather large, is wanted by another more
} influential person. It has been suggested that we attach a digital camera
} to our Hitachi H600 TEM to take over from the film camera we currently use.
} I have suggested that digital images are not yet able to produce images as
} good as our film can and that the digital image can not be increased to a
} much larger size without appreciable loss of resolution. Can any one please
} help us find a solution as this same influential person also has their beady
} little eyes on our laboratory!
} We need to know
} 1. How easy is it to digitise the TEM without losing the ability to
} still turn to film.
} 2. What sort of camera/system should we be looking at?
} 3. What sort of software is needed?
} 4. How much do these things cost?
}
} Sarah Ellis
}
} Trescowthick Research Centre
} Peter MacCallum Cancer Institute
} Locked Bag #1 A'Beckett Street
} Melbourne 8006 Victoria
} Australia
}
} Phone +61-3-9656 1244
} Fax +61-3-9656 1411
} Email s.ellis-at-pmci.unimelb.edu.au
}
}






From: Robert H. Olley :      R.H.Olley-at-reading.ac.uk
Date: Thu, 10 Dec 1998 15:05:12 +0000 (GMT)
Subject: SEM: Pedagogical

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have difficulty explaining to people how and when to use the AUTO FOCUS
and AUTO STIGMATOR controls on our Philips 515, and more important, when
not to. Although I have an instinctive understanding, I find it very hard
to put into words. Does anybody have any materials for public use that
might be helpful, and can explain to people the background involved?

+------------------------------------------------------------------------+
| Robert H.Olley Phone: |
| J.J.Thomson Physical Laboratory {direct line +44 (0) 118 9318572 |
| University of Reading {University internal extension 7867 |
| Whiteknights Fax +44 (0) 118 9750203 |
| Reading RG6 6AF Email: R.H.Olley-at-reading.ac.uk |
| England URL: http://www.reading.ac.uk/~spsolley |
+------------------------------------------------------------------------+






From: Sara Miller :      saram-at-acpub.duke.edu
Date: Thu, 10 Dec 1998 10:45:29 -0500 (EST)
Subject: Re: salary + negative staining

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There's also a whole book on negative staining:

Hayat MA, Miller SE. 1990. Negative Staining. McGraw-Hill, NY.

Disclaimer: I have a modicum of interest in this book.

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735








From: shAf :      mshaf-at-darkwing.uoregon.edu
Date: Thu, 10 Dec 1998 08:16:42 -0800
Subject: RE: Pedagogical

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Robert writes ...
}
}
}
} I have difficulty explaining to people how and when to use
} the AUTO FOCUS and AUTO STIGMATOR controls on our Philips 515,
} ...

I would explain their use in terms of "intelligence algorythms" ...
that is, I've seen "auto mode" work well with certain objects imaged,
and how well they work being dependent on symetries assumed and present.
Your users should know that for all objects however, there is no better
"intelligence algorythm" than the human eye working with the human
brain.

... hope this helps :o)

cheerios, shAf

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - ICQ 210524
Geological Science's Electron Probe Facility - University of Oregon
mshaf-at-darkwing.uoregon.edu - http://darkwing.uoregon.edu/~mshaf/








From: Jeannine Caesar :      caesar-at-retina.anatomy.upenn.edu
Date: Thu, 10 Dec 1998 11:25:24 -0500 (EST)
Subject: TEM - serial sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

My colleague and I are trying to cut serial sections, but are having a
hard time getting the sections to ribbon nicely. She has tried trimming
by hand with a razor and on the microtome with glass knives. If anyone
has had good luck with their sections ribboning and would be able to give
some helpful hints, we would really appreciate it. Thanks.

Jeannine Caesar
Department of Neuroscience
University of Pennsylvania






From: rschoonh-at-sph.unc.edu
Date: Thu, 10 Dec 1998 09:21:04 -0500 (Eastern Standard Time)
Subject: Re: ? mounting renal tubules for immuno

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Doug,

You might want to try the following, adhear it to a slice of the 'meaty'
portion of cucumber an then section in a cryostat. I did this for orientation
and sectioning of gastric biopsies 15 or so years ago.

-- Begin original message --

}
} I have an interesting technical problem that I'm hoping the folks on this
} list can help me figure out.
}
} I have a colleague who is trying to embed dissected renal tubules (20
} micron diameter x 500 micron length) for immunofluorescence. He would like
} to be able to "tack down" individual tubules to some kind of substrate, so
} that he can keep track of which end he's working from, and would like to
} cross section them (1-4 micron sections) from end-to-end (vs longitudinal
} sx). We've tried adhering tubules to Thermanox coverslips and making
} cryotome sections, but the sections of plastic curled up like a scroll.
} We've tried Aclar film, and plastic embedding using LR Gold and had the
} same problem. Is there someone out there who has a suggestion? We have 2
} concerns, 1) we don't want to heat the specimen above physiologic temps and
} 2) we want to keep track of orientation so as to have cross sections of the
} tubules - adherence to some sort of substrate is one possible solution -
} any alternatives?
}
} Thanks for your help.
} Doug
} ....................................................................
} : Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy :
} : Research Specialist, Principal University of Arizona :
} : (office: AHSC 4212A) P.O. Box 245044 :
} : (voice: 520-626-2824) Tucson, AZ 85724-5044 USA :
} : (FAX: 520-626-2097) (email: doug-cromey-at-ns.arizona.edu):
} :...................................................................:
} http://www.pharmacy.arizona.edu/exp_path.html
} Home of: "Microscopy and Imaging Resources on the WWW"
}
}
}

-- End original message --

regards,
Bob
Robert Schoonhoven
Laboratory of Molecular Carcinogenesis and Mutagenesis
Dept. of Environmental Sciences and Engineering
University of North Carolina
CB#7400
Chapel Hill, NC 27599
Phone
office 919-966-6343
Lab 919-966-6140
Fax 919-966-6123

**I'm willing to make the mistakes if someone else is willing to learn from
them**






From: William McManus :      billemac-at-bioserver.vsb.usu.edu
Date: Thu, 10 Dec 1998 10:26:58 -0700
Subject: RE: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



To all concerned:
I am also in the process of digitizing our tem. We have a Zeus 902 cem,
fortunately it has a side mount for a 35mm camera already in place. Even
with a very good camera, there must be a loss of resolution compared to
film. Electrons exposing the film emulsion are highly resolved, whereas a
fluorescent image is inherently less resolved. We have always said that you
can see more on the negative than the view screen. Has anyone evaluated
different resolution cameras to see at what point pixel array size becomes
irrelevant when imaging a fluorescent screen?

William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920







From: Michael Bode :      mb-at-soft-imaging.com
Date: Thu, 10 Dec 1998 10:43:36 -0700
Subject: RE: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sarah,

Since you raise questions that probably are of interest to many people,
I will try to provide some information. Please keep in mind, that my
company, Soft Imaging System Corp, produces and sells these systems and
I might therefore be a little "biased" towards digital imaging. I will
also forward your email to our agent responsible for Australia so he can
provide you with more specific information.

1. How easy is it to digitise the TEM without losing the ability to
still turn to film.

That is usually not a problem. The digital cameras are either
side-mounted on a 35 mm port or bottom mounted below the film chamber.
in most cases (if not all), you can retain the film camera and take
images on negatives if that is required.

2. What sort of camera/system should we be looking at?

This and the following questions require a bit more thought. The answer
usually depends on what you want to to. As I mentioned above, there are
basically two different types of camera: side-mounted and bottom
mounted. One clear differentiation is the resolution and field of view
you can get from these cameras. Since the side-mounted cameras are
mounted above the viewing chamber, they "see" about the same area as a
negative, sometimes more. On the other hand, the bottom mounted cameras
usually see only a small part of the negative area. The side-mounted
cameras are better suited for applications where field of view is an
issue, the bottom mounted ones are better suited for applications where
resolution is the most critical.

Let's have a quick look at side-mounted cameras. They come in various
types and configurations. The simplest type is a simple TV camera, where
the CCD chip is either lens or fiber coupled to a small scintillator
that intercepts the beam (removable). This type of camera is great for
teaching and finding areas on a sample, as you get a real-time image on
a screen, but the low resolution (640x480 for US standards) makes it
less desirable for actual image acquisition (but see below, Multiple
Image Alignment).

Another type of cameras are digital cameras for side-mount. They come in
many possible sizes and configurations. Usually they have a better
resolution and a higher dynamic range (12 bit as opposed to 8 bit for TV
cameras), but often they require specific frame grabbers, and the frame
rate (numbers of images per second) can drop significantly. That does
not seem to be to important at first, but try to adjust astigmatism with
a camera that delivers less than 10 frames per second. I personally find
it impossible to do. I would say, that a good camera must deliver a
resolution of at least 4 times TV (i.e., 1280 x 1024), have a dynamic
range of at least 12 bit and a frame rate of 10 or more at full
resolution. It should also be able to accept exposure times of several
seconds for those darkfield images, which proabably means cooling the
camera.

Then there are specialty cameras: high sensitivity cameras, extremely
fast cameras, etc. Those can be used, but they are normally not required
for "normal" TEM.

Now come the bottom mounted cameras. Almost all of them are fiber
coupled for better sensitivity. They are usually square and come in
1Kx1K and 2Kx2K chips. The 2K chips are MUCH more expensive than the 1K
chips. They ususally do not have the readout speed of side-mounted
cameras, and are basically replacements for negatives. Their lower read
out frequency allows better digitization and they are frequently
digitized to 14 bit or better. they do only see a part of the negative,
and together with the lower readout speed, it can be hard to find the
correct area. Most of them, however, have some features (binning, etc.)
that allows faster image repetition rate at a reduced resolution. these
cameras are often liquid cooled and either hook up to the TEM cooling
system or have their own chiller.

Of course you can have both cameras attached at the same time. That way
you can pick what camera to use depending on the application.

3. What sort of software is needed?

That again depends on what you want to do. As a minimum I would think,
that you need image acquisition from the camera you select (obviously),
some gray scale manipulations, non-destructive overlays for
documentation, an image database for archiving and retrieving (should be
network capable and compatible with other databases), printing
capabilities that include Microscopy specific things like scale bars,
true magnification, etc. The software should be able to maintain
calibration data for different camera setups, if possible communicate
with the TEM, and, again if possible, allow for motorized stage control.
Important for TEM is also online shading correction to compenstate for
fluctuations in the scintillator and illumination, and real-time
contrast maximization. This is especially important for 12 bit cameras,
as you may have to look for the "right" 8 bits to display and illuminate
the sample all the time. Auto-contrast cuts down on the dose. Depending
on how you work, a single screen system (with the operating system and
the images sharing one monitor) or a dual screen system (with the images
displayed on a separate monitor) is best. If you want to use the digital
system for astigmatism correction, you need a real-time FFT feature. If
you need particle detection Fourier filters, Programming, etc., this
should be available also, but is not really necessary for simple image
acquisition. You want to keep this option open, though, by selecting
software that offers an upgrade path.

4. How much do these things cost?

Ahh, the most important question. The answer is: From anywhere from
several thousand $US to hundreds of thousands of $US, depending on
manufacturer, quality, software, hardware, cameras, and your demands. I
know, this answer is not very satisfying, but it is just as impossible
to answer as the question: how much does a car cost. As I said, I will
forward your email to our person in Malaysia, and he can answer those
more specific questions.

Multiple Image alignment: This is a way to overcome resolution
limitations of cameras. One simply takes a series of images that are
displaced but overlapping, and the software reconstructs a larger image
by montaging the single images. Best done with a motorized stage and
controlled by software, but also possible by hand. This can give you
very high resolution images (e.g., 3Kx3K) from a low resolution camera.
The price you pay for this is that you have to take an image series
instead of a single image and the processing takes some time (seconds).


Whew, that turned out to be longer than I had intended. Aplogies to
anybody who thinks it is too long.

If you have further questions, please contact me through email or call
me. you can also check out our website at

http:\\www.soft-imaging.de

Finally: I do not claim that this is complete. There are many more
issued that need to be addressed, but I did not want to make it even
longer.

Michael

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: info-at-soft-imaging.com


*******************************************************
Disclaimer:
Soft Imaging System produces and sells image acquisition
and processing systems. We therefore have a vested
interest in some of the items mentioned above.
*******************************************************

}
} ----------
} From: Ellis, Sarah[SMTP:s.ellis-at-pmci.unimelb.edu.au]
} Sent: Wednesday, December 9, 1998 10:18 PM
} To: Microscopy Listserver
} Subject: Digitization of the TEM
}
} Hello all again,
}
}
} It seems that cost cutting and space saving have finally started attacking
} us! Our darkroom, which I admit is rather large, is wanted by another more
} influential person. It has been suggested that we attach a digital camera to
} our Hitachi H600 TEM to take over from the film camera we currently use. I
} have suggested that digital images are not yet able to produce images as good
} as our film can and that the digital image can not be increased to a much
} larger size without appreciable loss of resolution. Can any one please help
} us find a solution as this same influential person also has their beady
} little eyes on our laboratory!
} We need to know
} 1. How easy is it to digitise the TEM without losing the ability to still
} turn to film.
} 2. What sort of camera/system should we be looking at?
} 3. What sort of software is needed?
} 4. How much do these things cost?
}
} Sarah Ellis
}
} Trescowthick Research Centre
} Peter MacCallum Cancer Institute
} Locked Bag #1 A'Beckett Street
} Melbourne 8006 Victoria
} Australia
}
} Phone +61-3-9656 1244
} Fax +61-3-9656 1411
} Email s.ellis-at-pmci.unimelb.edu.au
}
}





From: Margaret Brannigan :      brannign-at-asrr.arsusda.gov
Date: Thu, 10 Dec 1998 13:07:55 -0500
Subject: TEM: need Reynolds PbCt recipe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Disaster struck our lab recently when the recipe for Reynolds PbCt got
tossed, someone made off with our EM book, and my hard drive got wiped.
All in one month! My attempt to recreate from memory failed, judging from
the disgusting sections I subsequently observed in the TEM. We usually
make up 50mLs at at time, please help!

Thanks,

Peggy Brannigan

EM Lab
Floral and Nursery Plants Research Unit, National Arboretum
USDA, Beltsville, MD. USA 20705

(301) 504-6097







From: Elaine Humphrey :      ech-at-unixg.ubc.ca
Date: Thu, 10 Dec 1998 09:58:31 -0800
Subject: Re: TEM - serial sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Hello all,
}
} My colleague and I are trying to cut serial sections, but are having a
} hard time getting the sections to ribbon nicely. She has tried trimming
} by hand with a razor and on the microtome with glass knives. If anyone
} has had good luck with their sections ribboning and would be able to give
} some helpful hints, we would really appreciate it. Thanks.
}
} Jeannine Caesar
} Department of Neuroscience
} University of Pennsylvania

Sections don't ribbon when the top and the bottom of the trapezium is not
smooth and parallel. Sharp edges (not raggedy) are important. If they are
not parallel, they don't connect to form a ribbon.
Elaine


Dr. Elaine Humphrey
Biosciences Electron Microscopy Facility
University of British Columbia
6270 University Blvd, mail-stop Botany
Vancouver, BC
CANADA, V6T 1Z4
Phone: 604-822-3354
FAX: 604-822-6089
e-mail: ech-at-unixg.ubc.ca







From: Tim Bruchman :      timbruc-at-azstarnet.com
Date: Thu, 10 Dec 1998 11:40:54 -0700
Subject: Re: Looking for Time Lapse Software Source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------82E464E000FDAFF4FEB219FF
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

The company I work for (Compix) has time-lapse image analysis software that
supports digital cameras (Hamamatsu, Spot, Photometrics, but sorry not
Polaroid yet!). It is relatively inexpensive ($2K) and easy to use. It is
being used mostly for DIC/GFP applications. E-mail me if you are interested
in more details or phone my office, 520-749-2070 or our main office
800-393-2526.

Tim Bruchman

Western Regional Sales Manager
Compix Inc. Imaging Systems

R-Brooks Corl wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
}
} Several months ago I was told about a software developer who has some
} relatively simple "time-lapse" imaging acquisition programs for
} digital cameras. I have since lost track of the name and source.
} Does anyone know who this is?



--------------82E464E000FDAFF4FEB219FF
Content-Type: text/x-vcard; charset=us-ascii; name="vcard.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for Tim Bruchman
Content-Disposition: attachment; filename="vcard.vcf"

begin: vcard
fn: Tim Bruchman
n: Bruchman;Tim
org: Compix Inc.
adr: 11934 E. Makohoh Trail;;;Tucson;AZ;85749;USA
email;internet: timbruc-at-azstarnet.com
title: Western Regional Manager
tel;work: (800) 393-2526
tel;fax: (520) 749-7999
tel;home: (520) 749-2070
x-mozilla-cpt: ;0
x-mozilla-html: FALSE
version: 2.1
end: vcard


--------------82E464E000FDAFF4FEB219FF--






From: Bruce Brinson :      brinson-at-rice.edu
Date: Thu, 10 Dec 1998 12:48:53 -0600
Subject: Digital(TEM)images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Sarah & friends,

In response to the original inquiry, I have worked with 2 imaging
systems. One uses an external camera (Kodak Mega +) focused on a
phosphor screen. It is mounted on the bottom of the column. The film
camera is available. It's ok for lower mag ( {x500K) work but suffers in
low light applications like HRTEM. It also suffers from spherical
aberration although most users don't notice. Can't really give you a
cost figure because it was a package deal. It is a good general use
system for our multi user facility. Cost aside though I would not
replace it with a similar system.
The other system I have used was Gatan model 794. This system was
quite impressive. The CCD is in the beam but can be retracted to allow
the use of film. Viewing area is about the same with both cameras (~1cm
dia). Due to cooling of the detector the Gatan camera has a couple of
issues that make me less comfortable putting it into a multi user
facility. I little interlocking could cure this. Depending on options
the 794 is in the 80-90K$ range.
Now all of that being said, while digital images will get the data.
When I make presentations to people sitting around large thick oak
tables. I want photos. Eye wash counts as much as data to some execs.
To scanners: if you keep a dark room & print 8x10s you can scan them
on a few hundred dollar flatbed scanner & do pretty well. I was suprised
at the results. Of course the files can be huge (but cheap).
As long a this subject is up. Has anyone had any experience with the
Juji FDL 5000 system. Seems it consist of a secret imaging plate that
can be taken from the instrument, electronically scanned by their
hardware, erased & re used. It's not clear from the literate if it
uses your existing camera or replaces your film camera but they are
claiming that you get the same viewing area. (real short coming of all
other digital systems I've seen)

Disclaimer: I have no business interest the mentioned companies. These
are just my thoughts today.

Bruce Brinson
Rice U.


When you travel at the speed of light, you don't need a review mirror.






From: Garry Burgess :      GBurgess-at-exchange.hsc.mb.ca
Date: Thu, 10 Dec 1998 13:23:23 -0600
Subject: Zeiss TEM Available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a Zeiss TEM, model 109 available immediately, if anyone out
there is interested. Please respond via e-mail.





From: Randy Tindall :      rtindell-at-NMSU.Edu
Date: Thu, 10 Dec 1998 12:41:03 -0700
Subject: Re: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm not completely up to date on the new digital technologies, but I concur
with the person who advised capturing images on film and scanning them in,
at least as far as the TEM is concerned. Film processing takes little
space, unless you're really shooting a lot, and scanners are a lot cheaper
than decent digital capture systems.

Also, I have yet to see a TEM digital capture system that can generate
really useful images, i.e. publication quality. As I said, however, things
may have changed.

SEM digital systems are much better, in my experience, and I assume this is
because they literally take over the signal capture and image generation
electronics of the scope. TEM's, of course, work completely differently.

Images on film are still the highest resolution, most archival, platform
independent, and cheapest form of data storage. I do believe, however,
that the printing side of photography can now be safely replaced by
reasonably priced scanners and printers. Even cheap inkjet printers now
are yielding surprisingly good images. These things and Adobe Photoshop
are quickly making printing darkrooms a thing of the past for the vast
majority of purposes.

Randy


Randy Tindall
Electron Microscope Laboratory
Box 3EML
New Mexico State University
Las Cruces, NM 88003

rtindell-at-nmsu.edu (work)
nrtindall-at-zianet.com (home)





From: Wilcox, Dan :      wilcoxd-at-amp.com
Date: Thu, 10 Dec 1998 15:31:40 -0500
Subject: LM: Zeiss deep UV microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello!

I am attempting to gather info on the Zeiss Axiotron 2 UV optical
microscope. I would appreciate hearing from any users concerning its
approximate cost, ease of use, and applications.

We are a semiconductor Fab making gallium arsnide 4 inch wafers. Our
application for the Axiotron is to determine if 0.5 micron wide optical
photoresist lines, which are about 1 micron deep, are fully exposed and
opened. This would be in a production environment, where our operators would
be inspecting hundreds of these wafers a week. We can successfully make this
determination using our Hitachi 4500 FESEM, but this is of course not
practical for production.

Please feel free to reply to:

Daniel Wilcox
Senior Eng.
M/A-COM Fab II
1-888-276-7975, x5847

email: wilcoxd-at-macom.com

Thanks!!!





From: MIKE ROCK :      merock-at-du.edu
Date: Thu, 10 Dec 1998 14:35:28 -0700 (MST)
Subject: Re: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sarah-
I have been through this discussion many times with current and former
employers.
First feel fortunate the bean counters are requesting the change. You will
find the cost of digital imaging will cost more than the current value of
your TEM. Try to get the slow scan CCD digital camera, the TV or video
rate cameras are not acceptable for image archiving.
These slow scan CCD cameras and their software usually cost $50-60,000 US.
Gatan and Dage both make nice systems. I'm sure there are others.
get the most for your money, buy the best technology, it will be obsolete
sooner than you wish. Working prints can be printed on a 300-1200 DPI
laser jet, and look fine, high quality photograde prints are usually
printed on a Dye-sublimation printer, kind of spendy, but again well worth
the invesment. And you feelings on film vs digital images are correct,
You can always go back and use TEM film for publication images if needed.
good luck & have fun shopping
-Mike

On Thu, 10 Dec 1998, Ellis, Sarah wrote:

} Hello all again,
}
}
} It seems that cost cutting and space saving have finally started attacking
} us! Our darkroom, which I admit is rather large, is wanted by another more
} influential person. It has been suggested that we attach a digital camera
} to our Hitachi H600 TEM to take over from the film camera we currently use.
} I have suggested that digital images are not yet able to produce images as
} good as our film can and that the digital image can not be increased to a
} much larger size without appreciable loss of resolution. Can any one please
} help us find a solution as this same influential person also has their beady
} little eyes on our laboratory!
} We need to know
} 1. How easy is it to digitise the TEM without losing the ability to
} still turn to film.
} 2. What sort of camera/system should we be looking at?
} 3. What sort of software is needed?
} 4. How much do these things cost?
}
} Sarah Ellis
}
} Trescowthick Research Centre
} Peter MacCallum Cancer Institute
} Locked Bag #1 A'Beckett Street
} Melbourne 8006 Victoria
} Australia
}
} Phone +61-3-9656 1244
} Fax +61-3-9656 1411
} Email s.ellis-at-pmci.unimelb.edu.au
}
}






From: MIKE ROCK :      merock-at-du.edu
Date: Thu, 10 Dec 1998 14:45:15 -0700 (MST)
Subject: Re: TEM - serial sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jeannine-
I remember an old trick using the glue adhesive from double
stick tape, solvated in acetone, then allow exess acetone to evaporate
off, once the appropriate consistancy is obtained use this "glue on the
bottom edge of the block face. ribbons are easy to cut, but sometimes
difficult to separate, don't cut too many at once.
-Mike
Thu, 10 Dec 1998, Jeannine Caesar wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hello all,
}
} My colleague and I are trying to cut serial sections, but are having a
} hard time getting the sections to ribbon nicely. She has tried trimming
} by hand with a razor and on the microtome with glass knives. If anyone
} has had good luck with their sections ribboning and would be able to give
} some helpful hints, we would really appreciate it. Thanks.
}
} Jeannine Caesar
} Department of Neuroscience
} University of Pennsylvania
}
}
}






From: Alan W. Nicholls :      nicholls-at-uic.edu
Date: Thu, 10 Dec 1998 15:38:14 -0600
Subject: Re: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sarah

We have a new em facility that was built without a darkroom. Our two main
TEM's both have slow scan CCD systems for acquisition and TV rate systems
for set up.

At 16:18 1998-12-10 +1100, you wrote:
} Hello all again,
}
}
} It seems that cost cutting and space saving have finally started attacking
} us! Our darkroom, which I admit is rather large, is wanted by another more
} influential person. It has been suggested that we attach a digital camera
} to our Hitachi H600 TEM to take over from the film camera we currently use.
} I have suggested that digital images are not yet able to produce images as
} good as our film can and that the digital image can not be increased to a
} much larger size without appreciable loss of resolution.

This is probably still true, just. Low magnification pictures are a
problem (8K on the microscope gives us a digital image of x80K. You have to
remember that the CCD sees what the small focusing screen looks at. At
high mags the maximum final useable image is around 10-20x the indicated
microscope magnification.

} Can any one please
} help us find a solution as this same influential person also has their beady
} little eyes on our laboratory!
} We need to know
} 1. How easy is it to digitise the TEM without losing the ability to
} still turn to film.

Easy. Our digital cameras bolt onto the bottom of the column or it is
possible to fit a camera side entry, above the viewing screen. In both
cases the film camera is still intact. The CCD camera's are cooled,
however, so more care has to be taken to ensure as good a vacuum as
possible in the camera chamber.

} 2. What sort of camera/system should we be looking at?

You need both a slow scan CCD and Intensified TV camera. We purchased ours
from Gatan.

} 3. What sort of software is needed?

We run with Digital Micrograph. The CCD control is integrated within it.

} 4. How much do these things cost?

Around $100K for a 1Kx1K CCD. The larger the CCD the better!

Regards


Alan W Nicholls, PhD
Research Electron Microscopist
Manager - Electron Microscopy Laboratory

Mailing Address:
Research Resources Center - East (M/C 337)
Room 100 Science and Engineering South Building
The University of Illinois at Chicago
845 West Taylor St
Chicago, IL 60607-7058

Tel: 312 996 1227
Fax: 312 996 8091
Office: Room 110





From: Lesley Weston :      lesley-at-interchange.ubc.ca
Date: Thu, 10 Dec 1998 14:15:37 -0800 (PST)
Subject: RE: the old boys network

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


When I started working as a junior histology tech in a hospital lab, many years
ago in England, I noticed that in all the hospital labs I knew of all junior
techs were female and all senior techs were male. I fled back to university
research labs before I could be forced to have the operation that seemed to be
required to climb the ladder.

Lesley Weston.



On Tue, 8 Dec 1998, Scott Ireland wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Let me put forth a 'possible' (i.e., no data to support this, just
} observation) explanation - as a vendor who has worked with the microscopy
} community for a few years, I have a noticed that many more women are
} entering this profession than men. I believe that I have read that this is
} true in most science (not necessarily engineering) fields today. If that
} observation is correct, then perhaps the lower average salary is a result of
} greater numbers of women new to the field.
}
} Yes, there does seem to be an old boys network at the top, but given the
} sheer numbers of women entering the field, that can't stay like that
} forever. I don't want to belittle this if it truly is a case of women doing
} the same job and being paid less - that certainly is not right - but I think
} that if my observations are correct, and have contributed to this sampling
} error - then maybe in the long run it is a good sign, and as these new
} microscopists gain experience the median salary will rise, and the
} differential will decrease.
}
} Just a thought, and only my own - not my employers -
}
}
} ------------------------------
} Scott Ireland
} North American Sales Manager
} Media Cybernetics, L.P.
} "The Imaging Experts"
} 716.473.0222 Tel
} 716.473.8048 Fax
} 888.691.2492 Pager
} scott-at-mediacy.com
} http://www.mediacy.com
} http://www.optimas.com
} -----------------------------
}
} -----Original Message-----
} } From: Dr. Rachel Teitelbaum [mailto:teitelba-at-aecom.yu.edu]
} Sent: Monday, December 07, 1998 5:26 PM
} To: Douglas W. Darnowski
} Cc: microscopy-at-sparc5.microscopy.com
} Subject: the old boys network
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} So how does one explain the dominance of men in the field, as evidenced by
} fewer tenured professors, etc., and I don't know if NIH breaks down its
} grants this way, but there are certainly many fewer Howard Hughes
} recipients who are female, and nobel laureates for that matter. I think
} it is a rational deduction that there is much less respect for women in
} science, and that despite the changing times, there is still an "old boys
} network" in place. Surveys in the business field have certainly
} demonstrated that women are often paid less despite similar
} qualifications. I'm sorry that more people did not respond to the survey
} to achieve statistical significance, as to how the results might have
} changed with more responses.....
}
} -Rachel
}
}
}
}
}






From: oshel-at-terracom.net (Philip Oshel)
Date: Thu, 10 Dec 1998 16:37:22 -0600
Subject: Re: SEM: Pedagogical

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Try using insects with lots of setae. These usually give auto controls
troubles, but can be dealt with manually. I've found dealing with these
structures helpful in showing when to use manual and how, and when to use
auto. Setae are also good subjects for training folks in the use of gamma
controls.

Phil

} I have difficulty explaining to people how and when to use the AUTO FOCUS
} and AUTO STIGMATOR controls on our Philips 515, and more important, when
} not to. Although I have an instinctive understanding, I find it very hard
} to put into words. Does anybody have any materials for public use that
} might be helpful, and can explain to people the background involved?
}
} +------------------------------------------------------------------------+
} | Robert H.Olley Phone: |
} | J.J.Thomson Physical Laboratory {direct line +44 (0) 118 9318572 |
} | University of Reading {University internal extension 7867 |
} | Whiteknights Fax +44 (0) 118 9750203 |
} | Reading RG6 6AF Email: R.H.Olley-at-reading.ac.uk |
} | England URL: http://www.reading.ac.uk/~spsolley |
} +------------------------------------------------------------------------+

****be famous! send in a tech tip or question***
Philip Oshel
Technical Editor, Microscopy Today
PO Box 620068
Middleton, WI 53562
(608) 833-2885
oshel-at-terracom.net








From: Wharton Sinkler :      wharton.sinkler-at-anlw.anl.gov
Date: Thu, 10 Dec 1998 17:23:14 -0600
Subject: Doyle-Turner potential parameters...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Dear Pedro,
There is a useful section on electron scattering factors in the
International Tables for X-Ray Crystallography. I don't know the volume
number, but it's not the one with the space group tables, and the section is
written by John Cowley.

While you won't find Doyle-Turner type parametrization of the scattering
factors in this section, the factors are tabulated as a function of
scattering angle, so you might be able to generate such parameters using a
fit.

I have a web page which may be useful for transforming from the
Doyle-Turner scattering factor parameters to equivalent parameters for the
potential. It is:
http://www.numis.nwu.edu/Staff/wharton/channeling.html

Good luck
Wharton
++++++++++++++++++++++++++++++++++++++++++++++++++
Argonne National Laboratory West
P. O. Box 2528
Idaho Falls, ID 83403
Tel: (208) 533-7724
FAX: (208) 533-7863

mailto:wharton.sinkler-at-anlw.anl.gov

----------
} From: Joao Pedro Esteves De Araujo {Joao.Pedro.Esteves.de.Araujo-at-cern.ch}
} To: sinkler-at-apollo.numis.nwu.edu
} Subject: Doyle-Turner potential parameters...
} Date: Thu, Dec 10, 1998, 8:27 AM
}

} Dear Sir
} I hope you can help me.
} I am trying to find the Doyle-turner potential parameters for
} Yttrium (Y). . Unfortunately in their original paper (Relativistic
} Hartree-Fock X-ray and Electron Scattering Factors, P.A. Doyle and P.S.
} Turner, Acta Cryst., A24, 390, 1968. ) these parameters were not
} calculated.
} Do you know some more recent reference where I could find this?
} Have you some suggestion how to find this information?
} I thank you in advance
} Yours Pedro
}
}
}
}
}












From: Tina Schwach :      schwa044-at-tc.umn.edu
Date: Thu, 10 Dec 1998 22:11:38
Subject: LV FESEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am writing a paper and I need some explanations of field emission guns
and why the resolution is so much better when using this gun in a
low-voltage SEM.

TIA
Tina








From: Tina Schwach :      schwa044-at-tc.umn.edu
Date: Thu, 10 Dec 1998 23:00:25 -0500
Subject: LV FESEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am writing a paper and need some basic information on how the
field-emission gun works and why the resolution is so much better with this
gun in a low-voltage SEM.
TIA
Tina Schwach







From: ROBIN CROSS :      R.Cross-at-ru.ac.za
Date: Fri, 11 Dec 1998 08:29:42 GMT+0200
Subject: Re: TEM: need Reynolds PbCt recipe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Margaret

} Disaster struck our lab recently when the recipe for Reynolds PbCt got
} tossed.

Here is the recipe (to make about 17ml):

Weigh out 0.44g lead nitrate and 0.59g of sodium citrate into a
small glass bottle. Add 10ml glass distilled water, securely fix the
cap and shake (by hand) violently for about 1 min. Transfer to a
mechanical shaker and allow to shake for a further 30 min. This
mixture should by then be white. To this mixture add 2.6ml of
freshly made-up 0.1M sodium hydroxide and gently agitate until the
cloudiness disappears, then add a further 4ml of distilled water.
NB. If any cloudiness persists discard and make it up again. Some
people use this diluted up to 1000x with 0.01M NaOH (the original
Reynolds reference has more details) but we use it as made up
above.

We store this stain in 3ml aliquots in Eppendorff tubes in the
refrigerator at about 4C and it keeps for ages.

I hope this puts you on track again.

Regards

Robin



Robin H Cross
Director : EM Unit, Rhodes University, Grahamstown, South Africa
eurc-at-giraffe.ru.ac.za - tel: +27 46 603 8168 - fax: +27 46 622 4377
http://www.ru.ac.za/affiliates/emu/em.htm





From: Laurent Ameye :      lameye-at-ulb.ac.be
Date: Fri, 11 Dec 1998 09:09:16 +0100
Subject: LVFESEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,

I am also interesting in the explanations of field emission guns.

TIA

Laurent







From: Reinhard Windoffer :      windoff-at-goofy.zdv.Uni-Mainz.de
Date: Fri, 11 Dec 1998 09:40:22 +0100 (MET)
Subject: microscopy cell chambers, FCS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello

We are looking for a cell-chamber to make time lapse recordings of cells.
Can someone tell me about experiences made with the FCS2 chamber from
Bioptechs? Since we want to make long time recordings focal stability is
of special interest for us.
If soemone has made expiriences with other systems I also would be
interested.

thanks
Reinhard

. . . . . . . . . . . . . . . . . . .
Dr. Reinhard Windoffer Fon: (00)49 (0)6131/39 3720
Universitaet Mainz Fax: (00)49 (0)6131/39 4615
Anatomisches Institut e-mail: windoff-at-mail.uni-mainz.de
Becherweg 13
D-55099 Mainz
Germany
. . . . . . . . . . . . . . . . . . .






From: NICOLA BOCK :      Nicola.Bock-at-nottingham.ac.uk
Date: Fri, 11 Dec 1998 10:06:19 GMT0BST
Subject: Reynold's lead citrate

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Margaret
I found my old book of recipes which includes Pb Cit.
It's a few years since I made any as I'm now in a materials em lab
but here's how we used to make 50ml:

1.33g lead nitrate
1.76g sodium citrate
30ml boiled distilled water

Mix in a 50ml volumetric flask, shaking vigorously for 1 min.
Leave to stand for 30mins with intermittant shaking.
Add 8ml of 1M sodium hydroxide (freshly mixed in boiled distilled
water).
Make up to 50ml with boiled distilled water and mix by inversion.

I hope this works.
Best wishes
Nikki


Nikki Bock
EM Technician
Dept. Materials Engineering
University of Nottingham
Nottingham NG7 2RD
(0115) 9513759/9513871
Email: emznjb-at-hermes.nottingham.ac.uk





From: Gary Dietrich Chinga :      garyc-at-stud.ntnu.no
Date: Fri, 11 Dec 1998 11:22:15 +0100 (MET)
Subject: Re: TEM - serial sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi

Try trimming the block with vertical leading face and sides; only the
trailing face keep the usual shape in order to avoid flexure of the block
during sectioning (This method is described by Fahrenbach 1984.
You can also use a solution of mounting media and toluene mixed
1:1 to coat the leading face facilitating the forming of the ribbon. This
procedure gave me very good results.

Gary.
NTNU-Trondheim
Norway.

On Thu, 10 Dec 1998, Elaine Humphrey wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} } Hello all,
} }
} } My colleague and I are trying to cut serial sections, but are having a
} } hard time getting the sections to ribbon nicely. She has tried trimming
} } by hand with a razor and on the microtome with glass knives. If anyone
} } has had good luck with their sections ribboning and would be able to give
} } some helpful hints, we would really appreciate it. Thanks.
} }
} } Jeannine Caesar
} } Department of Neuroscience
} } University of Pennsylvania
}
} Sections don't ribbon when the top and the bottom of the trapezium is not
} smooth and parallel. Sharp edges (not raggedy) are important. If they are
} not parallel, they don't connect to form a ribbon.
} Elaine
}
}
} Dr. Elaine Humphrey
} Biosciences Electron Microscopy Facility
} University of British Columbia
} 6270 University Blvd, mail-stop Botany
} Vancouver, BC
} CANADA, V6T 1Z4
} Phone: 604-822-3354
} FAX: 604-822-6089
} e-mail: ech-at-unixg.ubc.ca
}
}
}






From: Dan Hill :      dh2-at-mole.bio.cam.ac.uk
Date: Fri, 11 Dec 1998 10:29:52 +0000 (GMT)
Subject: Re: TEM: need Reynolds PbCt recipe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Margaret

The recipe for Reynold's lead stain is as follows:

Add 1.33g lead nitrate Pb(NO3)2 to 1.76g tri-sodium citrate Na3C6H5O7.2H2O
in a 50ml plastic (polymethylpentane) volumetric flask and add 30ml
boiled, cooled, CO2 free distilled water.
Invert the stoppered flask occasionally for 30 minutes and then add 8ml
CO2 free 1N NaOH and mix till clear.
Dilute to 50ml and leave to stand overnight before use.
Store in refrigerator.

Dan Hill
Department of Biochemistry
Cambridge University
UK

On Thu, 10 Dec 1998, Margaret Brannigan wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Disaster struck our lab recently when the recipe for Reynolds PbCt got
} tossed, someone made off with our EM book, and my hard drive got wiped.
} All in one month! My attempt to recreate from memory failed, judging from
} the disgusting sections I subsequently observed in the TEM. We usually
} make up 50mLs at at time, please help!
}
} Thanks,
}
} Peggy Brannigan
}
} EM Lab
} Floral and Nursery Plants Research Unit, National Arboretum
} USDA, Beltsville, MD. USA 20705
}
} (301) 504-6097
}
}
}
}






From: csedax-at-alpha.arcride.edu.ar
Date: Fri, 11 Dec 1998 08:12:14 -2359
Subject: SEM: problems while coating soft cheese samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,

we're having problems Au coating cheese samples for SEM observations.
They ared freezed and then coated. While being coated, after only a couple of
minutes, they get caramelized (does this word exists?). It looks like a film of
caramel is produced on the surface. Since we have now experience and no idea
about this type of materials we would like to hear any opinion about coating
conditions:
should we Au coat them at high pressure for a short time, or low pressure and
longer time? what makes this caramel appear?

Please, let us now if any of you have done something like this before. Thanks
in advance.

Silvia Montoro
Centro Regional de Investigacion y Desarrollo
Guemes 3450
3000 Santa Fe
Argentina
csedax-at-arcride.edu.ar





From: Dwight Beebe :      beebed-at-IRBV.UMontreal.CA
Date: Fri, 11 Dec 1998 07:42:06 -0400
Subject: Re: TEM digitization

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good morning:
We are also transitioning to digital with the following system: a
Microlumina slow scan, high-resolution CCD camera connected to a Mac G3, on
which we'll run Photoshop and aquire image from scans of negatives placed
on a special ($) high illumination output light box. The system is also
capable of scanning regular 2x2 and 120 slides. The camera will also be
used on my Olympus for brightfield work (not enough light to the camera to
permit fluorescence). We're looking at about 12-15 thousand for the whole
system.
I too have an H600 and have looked at cooled CCD cameras attached
to the 35mm port. As a previous respondent mentioned, they don't, as yet
offer the resolution that the Microlumina does and they cost about four
times as much. Hence our decision to go with scans of negatives. The
versatility of the system is also a plus. It is not in place yet, so I
don't know what the downside might be, but I think we'll be satisfied. I
also am reluctant to give up using the darkroom, although I have been very
happy with what can be accomplished using a digital darkroom, i.e.,
Photoshop.
Dwight

Dwight Beebe
Institut de recherche en biologie vegetale
Universite de Montreal
4101, rue Sherbrooke est
Montreal (Quebec) H1X 2B2 Canada
Tel: 514-872-4563
FAX: 514-872-9406







From: Rich Dudley :      rdudley+-at-pitt.edu
Date: Fri, 11 Dec 1998 08:26:16 -0500
Subject: Re: microscopy cell chambers, FCS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reinhard Windoffer wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hello
}
} We are looking for a cell-chamber to make time lapse recordings of cells.
} Can someone tell me about experiences made with the FCS2 chamber from
} Bioptechs? Since we want to make long time recordings focal stability is
} of special interest for us.
} If soemone has made expiriences with other systems I also would be
} interested.
}
} thanks
} Reinhard
}
} . . . . . . . . . . . . . . . . . . .
} Dr. Reinhard Windoffer Fon: (00)49 (0)6131/39 3720
} Universitaet Mainz Fax: (00)49 (0)6131/39 4615
} Anatomisches Institut e-mail: windoff-at-mail.uni-mainz.de
} Becherweg 13
} D-55099 Mainz
} Germany
} . . . . . . . . . . . . . . . . . . .

We use the FCS2 for our live cell imaging, and I think it is fantastic. It
is very well designed, and very easy to use. We use it pretty much as they
recommend, with their pumps and heaters. They have a web site
(www.bioptechs.com) that has a lot of useful info about live cell imaging, as
well as technical specs of their systems. It is not cheap, but worth it. My
one suggestion: get a couple extra of the thermoplastic glass and gaskets.
Careless people ruin these easily.

Your focal stability is going to depend a lot on your microscope system. If
you are using an immersion lens, the spring-loaded head will have a tendency,
over several hours, to push itself slightly out of focus. Gravity will also
play a small role in moving your objectives. The chamber itself is very
stable. Dan at Bioptechs is very helpful, and can give you much good advice.

rich

--
--- --- --- -- -- -- --- --- ---
Richard J. Dudley (rdudley+-at-pitt.edu)
Research Specialist V
Dept. of Cell Biology and Physiology
University of Pittsburgh
http://www.cbp.pitt.edu
---} search BIONET archives at http://www.bio.net {---






From: Mick Thomas :      mgt3-at-msc.cornell.edu
Date: Fri, 11 Dec 1998 09:26:22 -0800
Subject: repeated messages: salary survey

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

I have received multiple copies (at least four today) of a message sent on
Dec 8 re:salary survey thoughts. I am certain this is no fault of the
author but instead due to some computer system somewhere. Have others been
receiving multiple copies as well?

Mick Thomas
Mick Thomas
UHV-STEM Laboratory
E-1 Clark Hall
Cornell University
Ithaca, NY 14853

Phone: 607-255-0650
Fax: 607-255-7658
e-mail: mgt3-at-msc.cornell.edu





From: Wharton Sinkler :      wharton.sinkler-at-anlw.anl.gov
Date: Fri, 11 Dec 1998 08:00:52 -0600
Subject: Re: Doyle-Turner potential parameters...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry,

The correct URL for the web page mentioned below is:
http://www.numis.nwu.edu/internet/Staff/wharton/channel.html

Wharton

}
} Dear Pedro,
} There is a useful section on electron scattering factors in the
} International Tables for X-Ray Crystallography. I don't know the volume
} number, but it's not the one with the space group tables, and the section
is
} written by John Cowley.
}
} While you won't find Doyle-Turner type parametrization of the scattering
} factors in this section, the factors are tabulated as a function of
} scattering angle, so you might be able to generate such parameters using a
} fit.
}
} I have a web page which may be useful for transforming from the
} Doyle-Turner scattering factor parameters to equivalent parameters for the
} potential. It is:
} http://www.numis.nwu.edu/Staff/wharton/channeling.html
}
} Good luck
} Wharton
} ++++++++++++++++++++++++++++++++++++++++++++++++++
} Argonne National Laboratory West
} P. O. Box 2528
} Idaho Falls, ID 83403
} Tel: (208) 533-7724
} FAX: (208) 533-7863
}
} mailto:wharton.sinkler-at-anlw.anl.gov
}
} ----------
} } From: Joao Pedro Esteves De Araujo {Joao.Pedro.Esteves.de.Araujo-at-cern.ch}
} } To: sinkler-at-apollo.numis.nwu.edu
} } Subject: Doyle-Turner potential parameters...
} } Date: Thu, Dec 10, 1998, 8:27 AM
} }
}
} } Dear Sir
} } I hope you can help me.
} } I am trying to find the Doyle-turner potential parameters for
} } Yttrium (Y). . Unfortunately in their original paper (Relativistic
} } Hartree-Fock X-ray and Electron Scattering Factors, P.A. Doyle and P.S.
} } Turner, Acta Cryst., A24, 390, 1968. ) these parameters were not
} } calculated.
} } Do you know some more recent reference where I could find this?
} } Have you some suggestion how to find this information?
} } I thank you in advance
} } Yours Pedro
} }
} }
} }
} }
} }
}
}
}
}
}
}
}
}
}







From: William McManus :      billemac-at-bioserver.vsb.usu.edu
Date: Fri, 11 Dec 1998 08:32:43 -0700
Subject: SEM: problems while coating soft cheese samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Silvia,

We examine cheese routinely at USU. It sound as if the coating process is
allowing the cheese to warm up and the serum and/or the fat may be flowing
and carmalizing. Look at Food Structure Vol.12 (1993), pp.475-482 for
details on our protocol, If unavailable, let me know and I will get you a
copy.
Bill


William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920


-----Original Message-----
} From: "csedax-at-alpha.arcride.edu.ar"-at-Sparc5.Microscopy.Com
[mailto:"csedax-at-alpha.arcride.edu.ar"-at-Sparc5.Microscopy.Com]
Sent: Friday, December 11, 1998 1:12 AM
To: MICROSCOPY-at-Sparc5.Microscopy.Com


Dear all,

we're having problems Au coating cheese samples for SEM
observations.
They ared freezed and then coated. While being coated, after only a couple
of
minutes, they get caramelized (does this word exists?). It looks like a film
of
caramel is produced on the surface. Since we have now experience and no idea

about this type of materials we would like to hear any opinion about coating
conditions:
should we Au coat them at high pressure for a short time, or low pressure
and
longer time? what makes this caramel appear?

Please, let us now if any of you have done something like this before.
Thanks
in advance.

Silvia Montoro
Centro Regional de Investigacion y Desarrollo
Guemes 3450
3000 Santa Fe
Argentina
csedax-at-arcride.edu.ar





From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Fri, 11 Dec 1998 10:31:01 -0800
Subject: Re: ? mounting renal tubules for immuno

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} -- Begin original message --
}
} I have an interesting technical problem that I'm hoping the folks on this list can help me figure out.
} I have a colleague who is trying to embed dissected renal tubules (20
} } micron diameter x 500 micron length) for immunofluorescence. He would like to be able to "tack down" individual tubules to some kind of substrate, so that he can keep track of which end he's working from, and would like to cross section them (1-4 micron sections) from end-to-end (vs longitudinal sx). We've tried adhering tubules to Thermanox coverslips and making cryotome sections, but the sections of plastic curled up like a scroll. We've tried Aclar film, and plastic embedding using LR Gold and had the same problem. Is there someone out there who has a suggestion? We have 2 concerns, 1) we don't want to heat the specimen above physiologic temps and 2) we want to keep track of orientation so as to have cross sections of the tubules - adherence to some sort of substrate is one possible solution -
} } any alternatives?
} }
} } Thanks for your help.
} } Doug
} } ....................................................................
} } : Douglas W. Cromey, M.S. Dept. of Cell Biology & Anatomy :
} } : Research Specialist, Principal University of Arizona :
} } : (office: AHSC 4212A) P.O. Box 245044 :
} } : (voice: 520-626-2824) Tucson, AZ 85724-5044 USA :
} } : (FAX: 520-626-2097) (email: doug-cromey-at-ns.arizona.edu):
} } :...................................................................:
} } http://www.pharmacy.arizona.edu/exp_path.html
} } Home of: "Microscopy and Imaging Resources on the WWW"

Doug:

How about placing the tubules inside another tube such as a length of
small intestine or aorta? Adding a bit of gelatine might keep more than
one tubule from moving around. The enveloping tube could be fixed to
make it less flexible if desired. Just a thought.

Geoff
--
***************************************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane Piscataway, NJ 08854
voice: (732)-235-4583; fax -4029 e-mail: mcauliff-at-umdnj.edu
***************************************************************





From: William McManus :      billemac-at-bioserver.vsb.usu.edu
Date: Fri, 11 Dec 1998 08:40:35 -0700
Subject: SEM: problems while coating soft cheese samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Silvia,

We examine cheese routinely at USU. It sound as if the coating process is
allowing the cheese to warm up and the serum and/or the fat may be flowing
and carmalizing. Look at Food Structure Vol.12 (1993), pp.475-482 for
details on our protocol, If unavailable, let me know and I will get you a
copy.
Bill


William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920
-----Original Message-----
} From: "csedax-at-alpha.arcride.edu.ar"-at-Sparc5.Microscopy.Com
[mailto:"csedax-at-alpha.arcride.edu.ar"-at-Sparc5.Microscopy.Com]
Sent: Friday, December 11, 1998 1:12 AM
To: MICROSCOPY-at-Sparc5.Microscopy.Com


Dear all,

we're having problems Au coating cheese samples for SEM
observations.
They ared freezed and then coated. While being coated, after only a couple
of
minutes, they get caramelized (does this word exists?). It looks like a film
of
caramel is produced on the surface. Since we have now experience and no idea

about this type of materials we would like to hear any opinion about coating
conditions:
should we Au coat them at high pressure for a short time, or low pressure
and
longer time? what makes this caramel appear?

Please, let us now if any of you have done something like this before.
Thanks
in advance.

Silvia Montoro
Centro Regional de Investigacion y Desarrollo
Guemes 3450
3000 Santa Fe
Argentina
csedax-at-arcride.edu.ar





From: Paul Webster :      pwebster-at-mailhouse.hei.org
Date: 11 Dec 98 09:15:01 -0800
Subject: RE: TEM - serial sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
MSA listserver submission {Microscopy-at-Sparc5.Microscopy.Com}
X-Mailer: QuickMail Pro 1.5.3 (Mac)
X-Priority: 3
MIME-Version: 1.0
Reply-To: Paul Webster {pwebster-at-mailhouse.hei.org}
Content-Transfer-Encoding: quoted-printable
Content-Type: text/plain; charset="US-Ascii"



From: Paul Webster :      pwebster-at-mailhouse.hei.org
Date: 11 Dec 98 09:15:01 -0800
Subject: RE: TEM - serial sections

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Reply to: RE: TEM - serial sections
Dear Jeanine,

Cutting serial sections is very easy and Elaine Humphrey and Mike Rock =
have told you all you need to know. I can only add a couple of simple =
suggestions.

Make sure the top and bottom of the block are parallel with each other and =
with the knife edge. =

Shaping the sides of the block so that they are perpendicular to the knife =
will sometimes help form ribbons. =

Make sure there is an identifying characteristic in the section which can =
be used for orientation. This can be a slight nick in the side of the =
block or an easily identifiable part of the sample.

If possible, make the height o fthe block as small as possible. This will =
allow you to produce large numbers of sections before having to pick them =
up.

Mikes idea for using the sticky tape glue is good but as he points out, =
the glue is a little too effective in sticking the sections together. We =
apply a product called "Tacky Wax" to the bottom (and/or top) of the block =
and find it sticks the sections together but not too securely. As I do =
not know the supplier of this product I have been experimenting with =
substitutes. The best so far is the "Post-it" glue that is supplied on a =
stick from office suppliers. =

Regards,

Paul Webster


Jeannine Caesar wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America =

Paul Webster, Ph.D
House Ear Institute
2100 West Third Street
Los Angeles, CA 90057
phone:213 273 8026
fax: 213 413 6739
e-mail: pwebster-at-hei.org
http://www.hei.org/htm/apw.htm






From: Augusto_A_Morrone-at-notes.seagate.com
Date: Fri, 11 Dec 1998 10:50:59 -0600
Subject: TEM digital images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
id sma004522; Fri Dec 11 16:51:41 1998
Received: from sv-gw1.stsv.seagate.com (sv-gw1.stsv.seagate.com [134.204.14.95]) by auth1.seagate.com with SMTP id IAA06230 for {microscopy-at-sparc5.microscopy.com} ; Fri, 11 Dec 1998 08:51:32 -0800
Received: by sv-gw1.stsv.seagate.com(Lotus SMTP MTA Internal build v4.6.2 (651.2 6-10-1998)) id 882566D7.005C9023 ; Fri, 11 Dec 1998 08:51:02 -0800
X-Lotus-FromDomain: SEAGATE-at-INTERNET
To: microscopy-at-Sparc5.Microscopy.Com
Message-ID: {882566D7.005C8EB3.00-at-sv-gw1.stsv.seagate.com}


Sarah:

Although you have already received plenty of responses to your question on
"digitizing the TEM," it may be interesting to you and others concerned
with this issue to look at the article by Alwyn Eades in the current
"Microscopy Today", and articles by J. Brink ("Micron" Vol 27 (2), 1997,
129-139), and Sherman and Chiu ("Journal of Microscopy" Vol 188, 1997,
285-289) which deal with resolution of CCD cameras.

I am installing a TEM, and fortunately was able to get a darkroom along
with it. For the most part the CCD camera will do very well, with the
added advantage of transmitting images to the requester through the local
network. Measurements on digitized images obtained directly from the CCD
camera or from scanned negatives are conveniently done on the computer
monitor, and the whole process is dry and clean. For the highest
resolution work I expect to use negatives, and possibly scanned prints of
those negatives.

Augusto A. Morrone, Ph.D.
Senior Advisory Quality Engineer
Analysis Lab
Seagate Technology
7801 Computer Ave South
Bloomington, MN 55435-5489
Phone: (612) 844-5838
Fax: (612) 844-8247







From: Richard Mount :      rmount-at-sickkids.on.ca
Date: Fri, 11 Dec 1998 12:20:40 -0800
Subject: Capillary bed density

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Suggestions needed!

I have been asked to compare the relative density of capillary beds in
different regions of cortex (rodent). My current thinking is to use
cardiac perfusion (saline/formaldehyde) followed by perfusion with a dye
(which to use?) then cut either vibrotome or frozen sections. If the dye
does not leak beyond the bed, quantitative image analysis could be used
to determine %difference of dyed vs. undyed cortex.
Has anyone done this type of study? Any comments and/or suggestions are
welcome.

Thanks,

Richard Mount
Auditory Science Laboratory
Hospital for Sick Children
Toronto, Ontario CANADA





From: Barbara Foster :      mme-at-map.com
Date: Fri, 11 Dec 1998 12:22:31 -0500
Subject: Course announcement - not just for chemists!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Applied Optical Microscopy will be offered for the 15th year just prior
to PITTCON '99, in Orlando, Fl.


This certified American Chemical Society short course offers 3 days of
total immersion, hands-on experience in light microscopy. Although the
emphasis on materials analyses, the course offers very practical
information for microscopists from all disciplines. A key components of
the program are:

a) a special Saturday evening program on video and digital imaging

b) a full day on polarized light analysis (both quantitative and
qualitative)

c) the opportunity to bring your own samples and use them during the
course to develop practical strategies for solving microscopical
problems.


For further details and registration information, visit our website
{ {http://www.MME-Microscopy.com/education} or call me here at the
Microscopy/Microscopy office.


Barbara Foster

Course Coordinator




{color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
..Educating microscopists for greater productivity.

{/color} 125 Paridon Street Suite 102 Springfield, MA 01118

PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

Visit our web site { {http://www.MME-Microscopy.com/education}

******************************************************

{bigger} {bigger} MME {/bigger} {/bigger} is America's first national
consortium dedicated to customized on-site training in all areas of
microscopy, sample preparation, and image analysis {bigger} .

{/bigger}










From: Weiland, Hasso :      Hasso.Weiland-at-alcoa.com
Date: Fri, 11 Dec 1998 13:58:29 -0500
Subject: Vacuum impregnation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I need some advice in vacuum impregnation of fractures. We plan to
prepare cross sections for SEM analysis from partially fractured
aluminum samples (that is a sample which has a fine crack going though
its center). This analysis will be followed by opening the crack for an
analysis of the fracture surface itself. It is important to protect the
crack crevice from the attack by the solutions (water, etchants) used to
prepare the cross sections.

In our first attempt, we vacuum impregnated the crack with
stop-off-lacquer before handling. After opening the crack, fine
particles of the lacquer adhered to the rough fracture surface and we
were not able to remove these particles without attacking the fracture
surface as well.

Are there any waxes out there suitable for this application?

Many thanks
Hasso Weiland
Alcoa Technical Center
Alcoa Center, PA 15069

( 724 337-3133
Fax 724 337-2044
hasso.weiland-at-alcoa.com






From: Carol M. Garland :      cmg-at-cco.caltech.edu
Date: Fri, 11 Dec 1998 13:51:44 -0800
Subject: Re: TEM digitization

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

This is exactly what we did for similar reasons.

We are very happy with the results. It is amazing the "prints" you can
obtain using an Epson stylus 800 printer with photo quality paper and
Photoshop. Most of our students use this system exclusively now. Some of
us "old timers" still like the darkroom at times.

Our system is PC based, but feel free to contact me if you need.

Carol
}
}
} Good morning:
} We are also transitioning to digital with the following system: a
} Microlumina slow scan, high-resolution CCD camera connected to a Mac G3, on
} which we'll run Photoshop and aquire image from scans of negatives placed
} on a special ($) high illumination output light box. The system is also
} capable of scanning regular 2x2 and 120 slides. The camera will also be
} used on my Olympus for brightfield work (not enough light to the camera to
} permit fluorescence). We're looking at about 12-15 thousand for the whole
} system.
} I too have an H600 and have looked at cooled CCD cameras attached
} to the 35mm port. As a previous respondent mentioned, they don't, as yet
} offer the resolution that the Microlumina does and they cost about four
} times as much. Hence our decision to go with scans of negatives. The
} versatility of the system is also a plus. It is not in place yet, so I
} don't know what the downside might be, but I think we'll be satisfied. I
} also am reluctant to give up using the darkroom, although I have been very
} happy with what can be accomplished using a digital darkroom, i.e.,
} Photoshop.
} Dwight
}
} Dwight Beebe
} Institut de recherche en biologie vegetale
} Universite de Montreal
} 4101, rue Sherbrooke est
} Montreal (Quebec) H1X 2B2 Canada
} Tel: 514-872-4563
} FAX: 514-872-9406


Carol M. Garland, Member of the Professional Staff
MC138-78
California Institute of Technology
Pasadena, CA 91125

Tele:626-395-2168
Fax:626-795-6132
e-mail:cmg-at-cco.caltech.edu







From: Caroline Schooley :      schooley-at-mcn.org
Date: Fri, 11 Dec 1998 14:07:55 -0800 (PST)
Subject: Re: SEM: problems while coating soft cheese samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} we're having problems Au coating cheese samples for SEM
} observations.
} They are frozen and then coated. While being coated, after only a couple of
} minutes, they get caramelized (does this word exists?). It looks like a
} film of
} caramel is produced on the surface. Since we have no experience and no idea
} about this type of material we would like to hear any opinion about coating
} conditions:
} should we Au coat them at high pressure for a short time, or low pressure and
} longer time? what makes this caramel appear?
}
} Please, let us know if any of you have done something like this before. Thanks
} in advance.
}
} Silvia Montoro
} Centro Regional de Investigacion y Desarrollo
} Guemes 3450
} 3000 Santa Fe
} Argentina
} csedax-at-arcride.edu.ar

Sylvia -

You'll find the FOODS UNDER THE MICROSCOPE Web page
http://www.cyberus.ca/~scimat/ quite helpful. Contact its author (who
doesn't read this listserver) if you have further questions: Milos Kalab
{kalabm-at-EM.AGR.CA}


Caroline Schooley
Educational Outreach Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.MSA.microscopy.com/ProjectMICRO/Books.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html







From: Michael A. O'Keefe :      maokeefe-at-lbl.gov
Date: Fri, 11 Dec 1998 16:05:59 -0800
Subject: TEM chairs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------3E6C2B8D57EB3C4AA2B2B83D
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

All,

All the ergonomic chairs we can find contain (magnetic?) metal parts. If they
are moved at the microscope, their motion affects the position of the zero-loss
peak of a GIF. The effect can be rather annoying. We really need ergonomic but
non magnetic chairs made from plastic or wood. Does anyone out there know of a
supplier?

Dr. Christian Kisielowski

National Center for Electron Microscopy
Lawrence Berkeley National Laboratory
One Cyclotron Road, Bldg. 72/125
Berkeley, CA 94720
USA

Tel: (510) 486 4716
Fax: (510) 486 5888

--------------3E6C2B8D57EB3C4AA2B2B83D
Content-Type: text/x-vcard; charset=us-ascii; name="vcard.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for O'Keefe, Michael
Content-Disposition: attachment; filename="vcard.vcf"

begin: vcard
fn: Michael O'Keefe
n: O'Keefe;Michael
org: {center} {img src=http://mercproj.lbl.gov/logo.gif} {br} {b} {big} National Center for Electron Microscopy {/big} {/b} {/center}
email;internet: maokeefe-at-lbl.gov
title: Deputy Head
x-mozilla-cpt: ;0
x-mozilla-html: TRUE
version: 2.1
end: vcard


--------------3E6C2B8D57EB3C4AA2B2B83D--






From: Hoan Nguyen :      opea.hoan-at-wanadoo.fr
Date: Sat, 12 Dec 1998 10:05:43 +0100
Subject: Re: motorized table for SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,
Our company produced verry large chamber for SEM. You are invited to visi=
t our
site
web: http://perso.wanadoo.fr/opea.hoan/
Our realizations for:
CEAT (Centre d'Essais Aeronautique de Toulouse - FRANCE)
OCE (Holland)
Hoan
OPEA
114, rue de la Jarry
94300-VINCENNES (FRANCE)
Tel.33.1.4328 3496 Fax:33.1.4328.0364

Marienhoff a =E9crit:

} -----------------------------------------------------------------------=
-
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Co=
m
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.htm=
l
} -----------------------------------------------------------------------.
}
} Dear all,
}
} We want to integrate two separate motorized tables of five axis each in=
to our
} large chamber SEM. Who knows suppliers of these tables?
}
} Best regards
}
} Peter Marienhoff
} ___________________________
} Dr. Peter Marienhoff
} VisiTec Microtechnik GmbH
} Karl-Marx-Str. 14
} D-23936 Grevesmuehlen
} Germany
}
} Fon: +49-3881-790-47
} Fax: +49-3881-790-48
} email: pmarienhoff-at-visitec-em.de
} http://www.visitec-em.de









From: acoulter-at-greennet.net (Andrew Coulter)
Date: Sat, 12 Dec 1998 06:41:06 -0800
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: Louie Kerr :      lkerr-at-mbl.edu
Date: Sat, 12 Dec 1998 16:19:34 -0500
Subject: Re: microscopy cell chambers, FCS2

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Reinhard,

Two groups here at the MBL have been using the Delta TC3 system from
Bioptechs. It has been working very well for our purposes but the FCS2
would probably be better for time lapse recordings. Dan at Bioptechs has
been of much help along with their web page: www.bioptechs.com.

Thanks,
Louie

At 9:40 AM +0100 12/11/98, Reinhard Windoffer wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Louie Kerr
Research and Education Support Coordinator
Marine Biological Laboratory
7 MBL Street
Woods Hole, MA 02543
508-289-7273
508-540-6902 (FAX)

VISIT OUR WEB SITE:
http://www.mbl.edu







From: DVCCO-at-aol.com
Date: Sun, 13 Dec 1998 04:14:42 EST
Subject: CCD Digital 1K Color @ 12fpsRGB or color integration/New Site/Cell Bio Show

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


www.dvcco.com No cooling needed, color integration to 10 sec $4995 !!!
These are 10 bit cameras.... 60dB signal to noise not 50.
See model info below
This is an informational update on a new type digital camera on the market
from the manufacturer. This is current info, not some old, me too camera info
rehash from years past by someone who paid too much.
CURRENT SHOW: CELL BIOLOGY / SAN FRANCISCO MOSCONE BOOTH 15
12/13 - 12/16/ 1998 Going on as you read this!
Web site: www.dvcco.com this is the most current site for you to
bookmark
****** Please look at image gallery / color images/ biomedical images
HERE IS WHAT WE CAN NOW OFFER THAT THE OTHERS CANNOT FOCUSING ON THE
MICROSCOPY MARKET.
* A 12fps focusable image finally.... No 60% signal loss via a noisy RGB
sequential filter and slow response of 1 frame every 4 seconds etc. -at- $8k+.
* Full RS232 Control outboard of the camera for 10 second integration with no
noise
or up to 30dB gain at 12fps so you can have it either way. Remember if you
start with a 60dB signal and add gain you go from 10 bits down to 8 but if you
start with a competitors camera at 50dB which is 8 bits only marginally you go
down to 7 bits or 6. So with 50dB your are starting out with 7.5 to marginal
8 bits before you even add gain. Who wants less than 8 bits in the first
place???
Your camera guy might say Oh we use that sensor also BUT....what he doesn't
know is that due to his manufactures rush to get the camera to market they did
a sloppy auto route job of the board and ended up with 50dB signal to noise Vs
our 60dB. It is how much of the signal to noise of the sensor YOU PRESERVE
that counts in a design. Yes, tell your camera/microscope salesmen they can
stop telling everyone they need cooling just because of a little integration
in the range of 0-20 seconds. NOT SO if they have YOUR best interest in mind.
This was the big surprise at the Neuroscience show / virtually everyone
automatically parroted cooling in the same breath with integration....
* Color integration.... one snap and its done, no overlays.
* We offer the whole system digital camera, cable, supply, and frame grabber
PCI/PC by Epix / PIXCI-D for about $6300!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
$4995+200+150+995.
* Yes we have a new Mac interface also / ITI board, but please if you have a
choice go PC/ PCI. So what if NIH Image is free, so is this software....
We can even offer you the computer 400Mhz+ in a system package.
* We are a US camera manufacturer, with full control of our design and OEM
engineering capability. OEM quantity orders are no problem.
* Yes it is TWAIN compatible and the Epix board for $995 even has all the free
software you will ever need to grab and evaluate your image via some of the
other fine programs that are out there along with custom color balance
software for the DVC 1300C
* The DVC1300 Mono or DVC1300C / RGB camera can BOTH be USED on the same
RS422 digital only Mono Epix frame grabber. You do not need a special frame
grabber for color and the display of the RGB image is on your computer
monitor...not some expensive RGB monitor. Were talking digital here, not
analog so we have more avenues and flexibility than with the old analog
cameras.
* When evaluating cameras do it via S/N - signal to noise first....at 12
frames/sec
We offer 10 bits, no more... Even with integration. Be careful, others will
say they do 16 bits or 14 etc.... but at how much integration time...
* Yes we have cooling to -40 if needed. Most integration out to perhaps 10
seconds or a little more DO NOT. Cooling is an expensive add on that also
could reduce the life of the camera if the seal leaks later.....
* The sensor is effective a grade one sensor. Before you purchase a camera
ask what grade sensor you are getting. A well known mfgr ships grade 3
sensors and charges you 10K+ for the privilege.
Models:
DVC1300 Monochrome 1300 x 1030 pixels-at- 12fps or 1300 x 515 -at- 24fps
!!!!!!!!
DVC1300C Color RGB 1300 x 1030 pixels -at- 12fps

* Here is an idea:
For the price these systems are offered you can outfit your lab with 2 if not
more cameras with one frame grabber and 1 each of the RGB and mono DVC cameras
instead of some pricey mono camera with a noisy lossey sequential filter and
no frame grabber for the price, or some built in one with no flexibility.
We fit a niche market...Most people that would like a good fluorescent image
or other with decent signal to noise for 10 bit or less with gain picture and
all the gain 30db or integration they need. Yes, if you go past 20 or 30
seconds then you will need cooling. The explanation of the color filter/
Bayer pattern is on the web site under DVC1300C. The Bayer pattern reduces
aliasing also.
Every wonder why at the Vision or Photonics East/West shows you see these
color camera manufacturers showing big.......colorful teddy bear's etc. Its
because they do not have a decent Megapixel color camera, only low resolution
analog that is 7 bits or 48 dbS/N. Big and smooth does not show you the
detail.
There is a complete FAQ ....Frequently asked questions section in the index
for your reading pleasure. We hope this helps let you know there is some
radically new cost effective technology out there that you should be aware of.
Double click on the images to blow them up for your review and download.
Thank you for your time and if I offended anyone, God help me, I have seen
enough boring messages that I know this does not fall into.
Regards,
Rich

Richard Klotsche
DVC Company Sales Manager
619-444-8300
619-444-8321-fax
dvcco-at-aol.com
www.dvcco.com







From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Sun, 13 Dec 1998 10:30:38 -0500
Subject: SEM: problems while coating soft cheese samples

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

In the past I was deeply involved in the design of SEM cryo systems and t=
he
problem you relate is typical of what I would term "old fashioned" sputte=
r
coating systems, too hot!.

Sputter coating is one of the biggest problem areas in any form of SEM, i=
t
often cooks and very often decorates a specimen.

May I suggest you use the following procedure to look at the specimen
UNCOATED, then you will find the source of your problem.

1. Run at 2kV =

2. Depending upon the microscope you have you will probably need abo=
ut
50uA emission =

3. Use a spot size SMALLER than you would normally use as this is a
safety device
4. Run at a maximum of 15mm WD

Always look at a specimen in this way first and then you will know what t=
he
real specimen looks like. You will be surprised just how many specimens=

will be usable without coating. =


If you do decide to coat do so at the lowest voltage that your coater wil=
l
provide (probably about 800v) and run for no longer than 1 minute at 20mA=
. =

To improve the coating efficiency move the sample along the cold block,
backwards and forwards, as you coat.

For more help please feel free to contact me with details of your cryo
system and your microscope.

Steve Chapman

Senior Consultant E.M.
Protrain, 16 Hedgerley, Chinnor, Oxford OX9 4TN, England.
Tel & Fax 44 (0)1844 353161
Web Site - http://ourworld.compuserve.com/homepages/protrain
For Consultancy and Courses in Electron Microscopy World Wide





From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Sun, 13 Dec 1998 10:30:34 -0500
Subject: FESEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi

The electron source sets the limits of the SEM system. All that our
condenser lenses do is to reduce the size of the source by throwing
electrons away. A crude rule of thumb is the resolution attainable will =
be
10,000X smaller than the source size (50 micron source 50 Angstrom
resolution). The answer to the FEGSEM sucess is its source. About 1000=

times brighter than a W hairpin and measured in nm rather than micron
there are far more electrons available to us per unit area. This makes =
it
possible to form smaller probes with sufficient current to gereate a high=

level of signal. The FEG really comes into its own at low voltage for a
number of reasons.

1. The W hairpin system loses efficiency as you move away from the
design kV. The gun has to be designed with a certain minimum anode to
cathode distance (about 1mm for every 2kV) in order to prevent discharge =
at
the highest kV. As soon as you move away from the higest kV the system i=
s
no longer optimised and the gun becomes less efficient. You can improve
the gun performance by raising the anode and moving the filament forward.=
=

Even with these modifications the low kV performance still falls short
because of the increased level of aberrations in the system. Lower
operating lens currents and accelerating voltages mean that small
irregularities have a greater effect. Add to these problems the increase=
d
beam spread when it strikes the specimen at lower voltages and you see w=
hy
the resulting performance is pretty poor.

2. Using a FEG we do not have the problems with gun design and the
source is so much improved that even when using small spot sizes at the
specimen (compensating for beam spread in the specimen and that due to
aberrations) there is more than enough current to generate a good signal.=


Take a W hairpin system at 2kV and you will be doing well to obtain 20,00=
0X
with any quality, go to a good FEG SEM and 90,000X at 2kV is not such a
problem. The best FEG systems from my experience are the ones using a
double detecting system, here the choice of signals makes the operation o=
f
the instrument and an understanding of imaging information an operator's
dream, we can select the signal that we want.

Dual detector imaging often known as semi in lens imaging, relies upon th=
e
SE being attracted to a detector which is situated above the final lens
pole piece. Initially commercial instruments with two ET detectors used
the field from the final lens to "pull" SE into the upper detector. ISI
were first with their SS Series and then Hitachi with the S570 took this
route. In these instruments they allowed the specimen to be anywhere
between 100% out of lens, through to what we would consider to be a WD of=

-5mm. That is 5mm up inside the pole piece! Problems arose if you were
using magnetic materials, hence the development of a twin detector system=

by Hitachi which used a field coil to drag electrons up to the detector a=
nd
a pole piece design which retained the lens field within the pole piece. =


The twin detector system gives the microscopist the best of all worlds.

1. The upper detector provides an opportunity to sift out the BSE an=
d
obtain a pretty pure SE image. The high lens strength at very short WD
(~3mm) enables the instruments to reach very high resolution levels. The
down side of this is, as SE are effected by charge, this mode is more pro=
ne
to charge problems. However at {5kV in my experience we have very few
problems provided the operator knows about low damage techniques.

2. The lower detector offers the type of contrast which we all see i=
n
our conventional single detector SEM, SE+BSE. The up side is that the BS=
E
contribution to this image results in far fewer charge problems and may b=
e
a good compromise for the biologist.

3. Add a BSE detector to such a system, and you can move in any
direction "pure" SE or SE + BSE or pure BSE. Drop the kV and the BSE
becomes even more interesting as the volumes of material involved almost
mimic SE volumes. Not to be discounted for biological applications.

So there we are I hope this helps those who have no experience with the
new era, that of dual detector FEG SEM.

Steve Chapman

Senior Consultant E.M.
Protrain, 16 Hedgerley, Chinnor, Oxford OX9 4TN, England.
Tel & Fax 44 (0)1844 353161
Web Site - http://ourworld.compuserve.com/homepages/protrain
For Consultancy and Courses in Electron Microscopy World Wide





From: Ellis, Sarah :      s.ellis-at-pmci.unimelb.edu.au
Date: Mon, 14 Dec 1998 11:08:10 +1100
Subject: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This message is in MIME format. Since your mail reader does not understand
this format, some or all of this message may not be legible.

------_=_NextPart_000_01BE26F5.D85999B0
Content-Type: text/plain;
charset="iso-8859-1"

THANK-YOU to all who responded to my questions. I received many informative
and wonderful responses. I am currently writing a submission to suggest we
keep at least half our darkroom for the developement of negatives and for
printing the occasional negative BUT I am also suggesting that we purchase a
good quality scanner with associatied software and computer in compensation
for losing part of our darkroom. Digitizing the TEM appears to be a very
expensive alternative but I'll also furnish the 'Powerful Ones' with
information on this as well. I will post a summary of the information I
received from all the wonderful Micrscopy Listserver people as soon as
practical. Again,

Thank-you!!!

Sarah Ellis

Trescowthick Research Centre
Peter MacCallum Cancer Institute
Locked Bag #1 A'Beckett Street
Melbourne 8006 Victoria
Australia

Phone +61-3-9656 1244
Fax +61-3-9656 1411
Email s.ellis-at-pmci.unimelb.edu.au


------_=_NextPart_000_01BE26F5.D85999B0
Content-Type: application/ms-tnef
Content-Transfer-Encoding: base64
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------_=_NextPart_000_01BE26F5.D85999B0--





From: Rosemary A. Walsh :      rw9-at-psu.edu
Date: Sun, 13 Dec 1998 20:54:46 -0500
Subject: problems while coating soft cheese

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Silvia,
Are you trying to keep your sample frozen
or even cooled during the coating +imaging? I have a couple
of thoughts depending on the equipment you have---
1. if you are using a cryo-SEM, keep the sample small enough
to fit into a 2 mm rivet. Place a second rivet onto the
surface, freeze in LN slush, transfer to the specimen
preparation chamber and fracture at -196C, transfer
the sample onto the SEM stage, heat the stage ( -100C)
and allow the surface ice to sublime (etch) for 4 min.
Try imaging your sample "uncoated" at 2.5 or 5 kV.
If you are unhappy with the image, transfer the
"etched" sample back to the table in the prep chamber
and coat 30 sec. with Au. This worked on "soft serve
froz. yogurt" and some protein gels
2. if you are doing all of this at ambient temperatures, you
might be able to cool the table inside the sputter-
coater by pumping ice water through it. We managed
to do this with an an old ISI PS-2 coater.In this
case I would increase the distance from the target
to your sample as much as is possible. In that same
coater the target housing can be pulled up and the
table adjusted as well.
3. if you can use your sputter coater to evaporate C or find
a vacuum evaporator, you can maximize the distance
from the C thread or rods as you evap. for 1 or 2 sec
at 1 sec intervals and again try to use a heat sink
such as a copper block in liquid N
4. lastly, you might try some form of fixation:
--freeze-subsituting the sample in acetone +osmium at -80C.
If you do this last procedure--be prepared to follow
up with a dehydration series and critical point drying.
The sample should be more conductive...but I would still
try to coat it with Au or C.
--follow a procdure used by M. Kalab (I have to check the
reference) but I believe a series of 3mm dia. holes were
drilled into an aluminum stub, the "soft" dairy sample was
placed into each hole and covered with low melting pt. agar.
The entire stub plus samples were fixed at ambient T with
gluteraldehyde followed by osmium tetroxide, ethanol
dehydration and critical point drying. Before the samples
were sputter-coated,the surface of each was touched
with the sticky tape on a second aluminum stub and both
were then sputter- coated. I recently tried this with
whey protein gels but only used vapor fixation with osmium
before dehydrating and coating.
I'll send references asap.
Let me know what you found to work.
Rosemary







From: Akarologie :      mites-at-nasmus.co.za
Date: Mon, 14 Dec 1998 09:12:16 +0200
Subject: (Fwd)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



------- Forwarded Message Follows -------






From: Eli Amundsen :      eli.amundsen-at-zoo.uib.no
Date: Fri, 11 Dec 1998 19:38:01 +0100
Subject: Photographies of mites

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can anybody help me?
I am in urgent need of SEM photos, good drawings and/or colour plates of
the following genera or species:
Dermatophagoides pteronyssinus; facies and whole body
Larva and adult of Neotrombicula autumnalis; dorsal view
Tetranychus sp.; facies and dorsal view - and if possible, drawings or
photos of the stages in their mating ritual + photo of infested plant
Varroa jacobsoni; facies, dorsal and ventral view + photo of infested
bees
Scheloribates laevigatus as intermediate host of the tapeworm Moniezia
expansa + drawings of lifecycle of the tapeworm
Dirocheles phalaenodectes; photo + drawins of lifecycle in the tympanal
organ of a noctuid moth
Demodex folliculorum or D. brevis; single specimen + drawing of several
specimens in hair follicle
Sarcoptes scabei; dorsal and ventral view + drawing of lifecycle in the
human skin
Dorsal view of male and female Ixodes ricinus, and colour plate of
engorged animal. Also SEM photos of the chelicera.
Arrenurus conicus; male and female, dorsal view
Glycyphagus canestrinii; male and female, dorsal view
Pergamassus crassipes, male and female, ventral view
Analges clavipes, male and female, ventral view
Labidocarpus megalonyx, Geckobia loricata, Caelculus echinipes,
Carabodes elongatus, and Pelops phaenotus, all dorsal views
Tegeocranus sp.(dorsal view) and Belba sp. (lateral view) with nymphal
exuvial remains intact
Pelops acromius larva, tritonymph and adult, all dorsal views

The pictures are to be used in an exhibition at the Bergen Aquarium, and
we need them before January 5th. And they must be high quality.

If you can help me, please send me an e-mail (eli.amundsen-at-zoo.uib.no)
naming the pictures you can supply, and your price of delivery and I
will get back to you with details of payments and postal adress.

Thank you all in advance for helping me.

Regards, Eli Amundsen








From: Bellex International Corporation :      bicde-at-bellexinternational.com
Date: Mon, 14 Dec 1998 06:33:05
Subject: Re: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 12/11/98 3:20 PM
Subject: FWD: Capillary bed density

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


A very easy way to evaluate capillary beds is to use the presence of
endogenous peroxide in the red blood cells present in the capillaries. Remove
the tissue from the animal and immediately cool...do not wash out the
blood but immersion fix in 4% paraformaldehyde. Then simply react vibratomed
tissue sections with DAB and the resultant dark brown reaction product
makes it quite easy to trace the capillaries by light microscopy. The
reaction can be further enhanced by exposing the sections to
peroxidase-congugated IgG to increase the background staining of the blood plasma.

See the reference: Sherman, D. and W. Paull. (1985) "New Method for
Visualization of Vascular Networks in Nonperfused Fixed Tissues". Stain
Technology 60:2 p.89.

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057


--------------------------------------


Suggestions needed!

I have been asked to compare the relative density of capillary beds in
different regions of cortex (rodent). My current thinking is to use
cardiac perfusion (saline/formaldehyde) followed by perfusion with a dye
(which to use?) then cut either vibrotome or frozen sections. If the dye
does not leak beyond the bed, quantitative image analysis could be used
to determine %difference of dyed vs. undyed cortex.
Has anyone done this type of study? Any comments and/or suggestions are
welcome.

Thanks,

Richard Mount
Auditory Science Laboratory
Hospital for Sick Children
Toronto, Ontario CANADA



RFC822 header
-----------------------------------

RECEIVED: from SF_Database by POP_Mailbox_-1298719063 ; 11 DEC 98
19:10:33 UT
Received: from SPARC5.MICROSCOPY.COM by mailcenter.btny.purdue.edu
with SMTP (QuickMail Pro Server for MacOS 1.0.3); 11 DEC 98
19:10:29 UT
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com
(8.6.11/8.6.11) id LAA09782 for dist-Microscopy; Fri, 11 Dec 1998 11:08:10 -0600
Received: from no_more_spam.com (Sparc5 [206.69.208.10]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id LAA09771 for
"MicroscopyFilteredEmail-at-msa.microscopy.com"; Fri, 11 Dec 1998 11:07:38 -0600
Received: from resunix.sickkids.on.ca (firewall.sickkids.on.ca
[192.75.158.8]) by Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id LAA09760 for
{Microscopy-at-sparc5.microscopy.com} ; Fri, 11 Dec 1998 11:07:21 -0600
Received: from sickkids.on.ca ([142.20.37.249]) by
resunix.sickkids.on.ca (8.7.5/8.7.5) with ESMTP id MAA04885 for {Microscopy-at-MSA.Microscopy.Com} ;
Fri, 11 Dec 1998 12:18:59 -0500
Message-ID: {36717E98.9785C54F-at-sickkids.on.ca}
Date: Fri, 11 Dec 1998 12:20:40 -0800
From: Richard Mount {rmount-at-sickkids.on.ca}
X-Mailer: Mozilla 4.5 [en] (Win95; U)
X-Accept-Language: en
MIME-Version: 1.0
To: Microscopy listserver {Microscopy-at-Sparc5.Microscopy.Com}
Subject: Capillary bed density
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit
Errors-to: Microscopy-request-at-sparc5.microscopy.com








From: Woody.N.White-at-mcdermott.com
Date: 12/11/98 12:58 PM
Subject: Vacuum impregnation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by gate.mcdermott.com (8.8.7/8.8.7) with ESMTP id JAA25382;
Mon, 14 Dec 1998 09:14:25 -0600 (CST)
Received: (from x400-at-localhost) by mta.mcdermott.com (8.7.1/8.7.1) id JAA16622; Mon, 14 Dec 1998 09:14:28 -0600 (CST)
X-Authentication-Warning: mta.mcdermott.com: x400 set sender to Woody.N.White%650-at-wtgw.mcdermott.com using -f
Received: by MCI; Mon, 14 Dec 1998 8:04:00 -0600


If chemistry associated with cracking is of concern (sounds that way), then
almost anything added to the crack will be problematic. Either it will add
it's
own chemistry, or you may clean off offending material along with the
impregnating material. If it is at all possible, I would suggest sectioning

(careful, dry cutting) the crack into two pieces, one for x-sect work and
the
other to open.

Woody White
McDermott Technology, Inc

______________________________ Reply Separator
_________________________________


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


I need some advice in vacuum impregnation of fractures. We plan to
prepare cross sections for SEM analysis from partially fractured
aluminum samples (that is a sample which has a fine crack going though
its center). This analysis will be followed by opening the crack for an
analysis of the fracture surface itself. It is important to protect the
crack crevice from the attack by the solutions (water, etchants) used to
prepare the cross sections.

In our first attempt, we vacuum impregnated the crack with
stop-off-lacquer before handling. After opening the crack, fine
particles of the lacquer adhered to the rough fracture surface and we
were not able to remove these particles without attacking the fracture
surface as well.

Are there any waxes out there suitable for this application?

Many thanks
Hasso Weiland
Alcoa Technical Center





From: mayte caldes :      maite.caldes-at-cnrs-imn.fr
Date: Mon, 14 Dec 1998 17:24:25 +0100 (MET)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by naimn3.cnrs-imn.fr (8.9.1a/jtpda-5.3.1-IMN) with ESMTP id RAA29984
for {microscopy-at-sparc5.microscopy.com} ; Mon, 14 Dec 1998 17:24:25 +0100 (MET)


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Unsubscribe

--
************************************
Maite Caldes-Rouillon
Institut des Matériaux de Nantes
2 rue de la Houssiničre, B.P. 32222
44322 Nantes Cedex 3
http://www.cnrs-imn.fr
Tél: +(33) 02 40 37 64 17
Fax +(33) 02 40 37 64 18
************************************







From: oshel-at-terracom.net (Philip Oshel)
Date: Mon, 14 Dec 1998 10:58:10 -0600
Subject: Re: Capillary bed density

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Richard,

Since you are doing relative densities, and don't need absolute measures
such as total cross-sectional area how about this: do corrosion-casting of
the brains--I don't have the references handy, other than Fred Hossler's
article in the Sept. '98 Microscopy Today. (Caveat--I'm tech. editor.)

Basically, methacrylate resin is injected into the blood system, then the
organ of interest is removed after the resin has polymerized. The brain in
your case. Then remove the areas of interest from the brains and digest
away the tissue. If you've removed equal sized cubes (etc.) from the
different areas/treatments, and if you've completely filled the
capillaries, etc., then the different weights of the casts would directly
correlate with the relative differences in capillary densities. Volume of
the capillaries could be determined by immersion in water--the small amount
of displace being the major headache. Alternatively, given the weight and
the known density of the resin, volume could be calculated.

The obvious problem is cutting away the resin casts of the capillary beds
from the casts of the larger vessels. This is possible, if mind-numbingly
delicate work.

Phil

} Suggestions needed!
}
} I have been asked to compare the relative density of capillary beds in
} different regions of cortex (rodent). My current thinking is to use
} cardiac perfusion (saline/formaldehyde) followed by perfusion with a dye
} (which to use?) then cut either vibrotome or frozen sections. If the dye
} does not leak beyond the bed, quantitative image analysis could be used
} to determine %difference of dyed vs. undyed cortex.
} Has anyone done this type of study? Any comments and/or suggestions are
} welcome.
}
} Thanks,
}
} Richard Mount
} Auditory Science Laboratory
} Hospital for Sick Children
} Toronto, Ontario CANADA

}}}}}}}}}}}}}}}}}}}}}}}}}}{{{{{{{{{{{{{{{{{{{{{{{{{{
Philip Oshel
PO Box 620068
Middleton, WI 53562
oshel-at-terracom.net








From: Ani issaian :      naz-at-cheme.caltech.edu
Date: Mon, 14 Dec 1998 09:19:33 -0800
Subject: unsuscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


please unsuscribe me until january 11, 1999.
Thanks,



Ani
--




Ani M Issaian
California Institute of Technology
Pasadena, CA. 91125
MC 210-41







From: Matt' Mail :      sales-at-electroimage.com
Date: Mon, 14 Dec 1998 18:20:26 -0500
Subject: Digitization of the TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.

------=_NextPart_000_0051_01BE278E.6BDDF460
Content-Type: text/plain;
charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

To all concerned:


Our company, ElectroImage, Inc. produces and markets the TEMSCAN System =
to digitize TEM negatives. =20

Whenever there is a thread about the various methods to obtain digital =
images from a TEM, our TEMSCAN system is generally mentioned by one of =
our customers as an alternative. What is not mentioned is what =
separates this system from flatbed scanners that have a transparency =
adapter. =20

The TEMSCAN System is composed of a MicroLumina digital camera, =
high-frequency light box, and copy stand. The result of this system is =
a 9MB grayscale file that is 2700 x 3400 pixels. Scan times are less =
than a minute for typical density TEM and SEM negatives. If resolution =
isn't as important, the scan times can be decreased by scanning with =
less resolution. Although many flatbed scanners are able to scan =
transparencies, the key element that sets the TEMSCAN system apart is =
the lens. With a lens attached to a scanning camera, cropping into the =
negative doesn't result in a loss of resolution. For example, if the =
area you wanted to scan on the negative was only one-fourth of the whole =
image, and both the scanner and the camera were capable of the same =
resolution (1000 x 2000 for example), the flatbed scanner would only =
give a 500 x 1000 pixel image while the camera can simply focus =
closer and still produce the full 1000 x 2000 resolution. Since most =
flatbed scanners are much larger than 4 x 5 inches (typically 8 x 10 =
or larger) you automatically throw away two thirds or the total =
potential resolution of the device. =20

The MicroLumina can also be used for brightfield microscope work =
producing a file of 26.1MB.

I would welcome the opportunity to scan in any negatives you wish to =
send me. I will return your negatives and give you a CD with your =
scanned images and hard copy to show the quality of this system.

Matt Irwin=20

ElectroImage, Inc.
277 Northern Blvd.
Suite 101
Great Neck, NY 11021
Phone: 516-773-4305
Fax: 516-773-2955
E-mail: sales-at-electroimage.com
Website: http://www.electroimage.com


=20

------=_NextPart_000_0051_01BE278E.6BDDF460
Content-Type: text/html;
charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

{!DOCTYPE HTML PUBLIC "-//W3C//DTD W3 HTML//EN"}
{HTML}
{HEAD}

{META content=3Dtext/html;charset=3Diso-8859-1 =
http-equiv=3DContent-Type}
{META content=3D'"MSHTML 4.72.3110.7"' name=3DGENERATOR}
{/HEAD}
{BODY bgColor=3D#ffffff}
{DIV} {FONT color=3D#000000 size=3D2} To all concerned: {BR} {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Our company, ElectroImage, Inc. =
produces and=20
markets the TEMSCAN System to digitize TEM negatives.  =
{/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Whenever there is a thread about the =
various=20
methods to obtain  digital images from a TEM, our TEMSCAN system is =

generally mentioned by one of our customers as an alternative.  =
What is not=20
mentioned is what separates this system from flatbed scanners that have =
a=20
transparency adapter.   {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} The TEMSCAN System is composed of a =
MicroLumina=20
digital camera, high-frequency light box, and copy stand.  The =
result of=20
this system is a 9MB grayscale file that is 2700 x 3400 pixels.  =
Scan times=20
are less than a minute for typical density TEM and SEM negatives.  =
If=20
resolution isn't as important, the scan times can be decreased by =
scanning with=20
less resolution.  Although many flatbed scanners are able to scan=20
transparencies, the key element that sets the TEMSCAN system apart is =
the=20
lens.  With a lens attached to a scanning camera, cropping into the =

negative doesn't result in a loss of resolution.  For example, if =
the area=20
you wanted to scan on the negative was only one-fourth of the whole =
image, and=20
both the scanner and the camera were capable of the same resolution =
(1000 x 2000=20
for example), the flatbed scanner would only give a 500 x 1000 pixel =
image while=20
the camera can    simply focus closer and still produce =
the full=20
1000 x 2000 resolution.  Since most flatbed scanners are much =
larger=20
than     4 x 5 inches (typically 8 x 10 or larger) =
you=20
automatically throw away two thirds or the total potential resolution of =
the=20
device.   {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} The MicroLumina can also be used for =
brightfield=20
microscope work producing a file of 26.1MB. {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} I would welcome the opportunity to =
scan in any=20
negatives you wish to send me. I will return your negatives and give you =
a CD=20
with your scanned images and hard copy to show the quality of this=20
system. {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT size=3D2} Matt Irwin {/FONT} {/DIV}
{DIV}   {/DIV}
{DIV} {FONT size=3D2} ElectroImage, Inc. {/FONT} {/DIV}
{DIV} {FONT size=3D2} 277 Northern Blvd. {/FONT} {/DIV}
{DIV} {FONT size=3D2} Suite 101 {/FONT} {/DIV}
{DIV} {FONT size=3D2} Great Neck, NY  11021 {/FONT} {/DIV}
{DIV} {FONT size=3D2} Phone:       =20
516-773-4305 {/FONT} {/DIV}
{DIV} {FONT=20
size=3D2} Fax:          =
 =20
516-773-2955 {/FONT} {/DIV}
{DIV} {FONT size=3D2} E-mail:        {A =

href=3D"mailto:sales-at-electroimage.com"} sales-at-electroimage.com {/A} {/FONT} {=
/DIV}
{DIV} {FONT size=3D2} {/FONT} {FONT color=3D#000000=20
size=3D2} Website:     {A=20
href=3D"http://www.electroimage.com"} http://www.electroimage.com {/A} {/FON=
T} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000=20
size=3D2}      {/FONT} {/DIV} {/BODY} {/HTML}

------=_NextPart_000_0051_01BE278E.6BDDF460--






From: Ellis, Sarah :      s.ellis-at-pmci.unimelb.edu.au
Date: Tue, 15 Dec 1998 11:05:15 +1100
Subject: Summary of TEM digitization

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This message is in MIME format. Since your mail reader does not understand
this format, some or all of this message may not be legible.

------_=_NextPart_000_01BE27BE.97C2BA30
Content-Type: text/plain;
charset="iso-8859-1"

As most of the answers to my queries on digitization of the TEM were posted
to the Microscopy Listserver, you may feel that this summary is not needed.
However, due to the large response generated, I feel the topic is worth
summarising. Please note that this summary is my own interpretation of the
overall views expressed and may not therefore be entirely accurate. Most of
the responses received were related to imaging for biologicaI sciences. I
have no affiliation with any commercial company.
It appears that the side mounted CCD cameras are the most popular mainly due
to the very high cost of the larger CCD cameras which are fitted below the
screen. These latter cameras also take up valuable knee space but if you
have lots of $$$ then it appears that the camera of choice here is a 2K x 2k
CCD camera. TV rate cameras or video cameras are really not suitable as the
images generated are of poor quality.
The general consensus is that the camera of choice in todays market is a
1024 x 1024 CCD camera with a dynamic range of at least 12 bit and a frame
rate of 10 or more at full resolution. The cost of such a camera appears to
be in the vicinity of US$50K -US$70K. The company 'Gatan' was frequently
mentioned as 'good'. Other companys mentioned were 'Soft Imaging System
Corp.' and 'Dage'. The company of choice will advise on whether you need
any other accessories depending on your needs and I did not receive enough
information on this to make a summary.
Software for these cameras is generally provided by the camera manufacturer
and it is wise to ensure that the software is not a proprietary image format
and is compatible with other software in use.
Having said all that, I found the overwhelming response from nearly every
reply was that the 'old fashioned' negatives were still superior in terms of
resolution compared with the digital 'equivalent'. The respondents that
were lucky enough to have a CCD camera attached to their TEM generally still
relied on film for publication quality prints and used the digital images
for 'working' prints.
In conclusion, unless you have lots of dollars, the way to go appears to be
to scan your TEM negatives using a good quality film scanner and manipulate
them with the associated software (Adobe photoshop). These scanned images
have the advantage that they can be easily cropped, annotated and assembled
into plates and can produce publication quality prints. In addition,
negatives are still the most archival, platform independent and cheapest
form of data storage.


I hope this summary is of some use,

Cheers

Sarah Ellis

Trescowthick Research Centre
Peter MacCallum Cancer Institute
Locked Bag #1 A'Beckett Street
Melbourne 8006 Victoria
Australia

Phone +61-3-9656 1244
Fax +61-3-9656 1411
Email s.ellis-at-pmci.unimelb.edu.au


------_=_NextPart_000_01BE27BE.97C2BA30
Content-Type: application/ms-tnef
Content-Transfer-Encoding: base64
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------_=_NextPart_000_01BE27BE.97C2BA30--





From: Corvos-at-aol.com
Date: Mon, 14 Dec 1998 20:18:12 EST
Subject: ISI SR50

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


LUC,

Can you confirm again that Santovac 5 is OK to use on the gun...

Regards,

Walter Protheroe





From: Gary M. Easton :      gary.easton-at-scannerscorp.com
Date: Tue, 15 Dec 1998 00:45:13 -0500
Subject: Older SEM needs a good home!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.

------=_NextPart_000_0019_01BE27C4.2D1E8240
Content-Type: text/plain;
charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

Attention listservers:
One of my government customers has decided to part with their ETEC =
SEM. They have asked me to find it a good home before it goes out on =
surplus. The recipient MUST be associated with the US government. The =
instrument is in excellent condition and has been under full service =
contract for at least 20(yes, that's TWENTY) years. It is has been =
upgraded with the following options:

Computerized mag control with text writer
Digital Scan Generator
DIGISEM SEM Digital Imaging package
Dapple(Noran) PC based EDS Analyzer(KEVEX Be detector needs to be =
rebuilt)
GW BSE
NEW 8 CFM inline roughing pump
Fairly new NESLAB water recirculator/chillier

For those of you who are not familiar with the ETEC SEM's, it's standard =
features are:

2.5 to 30KV - Tungsten filament
Smallish specimen chamber
Five axis eucentric specimen stage - Tilts to +45 degrees
Twin viewing CRT's
2000 line Photo CRT
Fast pump down time - approx. one minute from atmosphere to beam turn on

Remember, you have to be associated with the US gov't to get this =
instrument.

Please email me at: gary.easton-at-scannerscorp.com

Gary M. Easton, Pres.
Scanners Corporation
1-800-466-SCAN
Third Party SEM Service
EDS and Digital Imaging upgrades


------=_NextPart_000_0019_01BE27C4.2D1E8240
Content-Type: text/html;
charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable

{!DOCTYPE HTML PUBLIC "-//W3C//DTD W3 HTML//EN"}
{HTML}
{HEAD}

{META content=3Dtext/html;charset=3Diso-8859-1 =
http-equiv=3DContent-Type}
{META content=3D'"MSHTML 4.72.3110.7"' name=3DGENERATOR}
{/HEAD}
{BODY bgColor=3D#ffffff}
{DIV} {FONT color=3D#000000 size=3D2} Attention listservers: {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2}    One of my government =
customers has=20
decided to part with their ETEC SEM.  They have asked me to find it =
a good=20
home before it goes out on surplus.  The recipient=20
{STRONG} {EM} {U} MUST {/U} {/EM} {/STRONG} be associated with the US=20
government.  The instrument is in excellent condition and has been =
under=20
full service contract for at least 20(yes, that's TWENTY) years.  =
It is has=20
been upgraded with the following options: {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Computerized mag control with text=20
writer {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Digital Scan Generator {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} DIGISEM SEM Digital Imaging =
package {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Dapple(Noran) PC based EDS =
Analyzer(KEVEX Be=20
detector needs to be rebuilt) {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} GW BSE {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} NEW 8 CFM inline roughing =
pump {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Fairly new NESLAB water=20
recirculator/chillier {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT size=3D2} For those of you who are not familiar with the ETEC =
SEM's,=20
it's standard features are: {/FONT} {/DIV}
{DIV} {FONT size=3D2} {/FONT}   {/DIV}
{DIV} {FONT size=3D2} 2.5 to 30KV - Tungsten filament {/FONT} {/DIV}
{DIV} {FONT size=3D2} Smallish specimen chamber {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Five axis eucentric specimen stage - =
Tilts to=20
+45 degrees {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Twin viewing CRT's {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} 2000 line Photo CRT {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Fast pump down time - approx. one =
minute from=20
atmosphere to beam turn on {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {FONT=20
color=3D#000000} {STRONG} {EM} {U} Remember {/FONT} , you have to be =
associated with the=20
US gov't  to get this instrument. {/FONT} {/U} {/EM} {/STRONG} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT}   {/DIV}
{DIV} {FONT size=3D2} Please email me at:  {A=20
href=3D"mailto:gary.easton-at-scannerscorp.com"} gary.easton-at-scannerscorp.com=
{/A} {/FONT} {/DIV}
{DIV} {FONT size=3D2} {/FONT}   {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} Gary M. Easton, Pres. {/FONT} {/DIV}
{DIV} {FONT color=3D#000000 size=3D2} {/FONT} {FONT size=3D2} Scanners=20
Corporation {/FONT} {/DIV}
{DIV} {FONT size=3D2} 1-800-466-SCAN {/FONT} {/DIV}
{DIV} {FONT size=3D2} Third Party SEM Service {/FONT} {/DIV}
{DIV} {FONT size=3D2} EDS and Digital Imaging upgrades {/FONT} {/DIV}
{DIV} {FONT size=3D2} {/FONT}   {/DIV} {/BODY} {/HTML}

------=_NextPart_000_0019_01BE27C4.2D1E8240--






From: Ronnie Houston :      rhh1-at-airmail.net
Date: Tue, 15 Dec 1998 00:24:15 -0800
Subject: Cryoultramicrotomy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Need some help here. We are new to the cryo side of ultramicrotomy.
I would be interested in the clinical research applications of
cryoultramicrotomy.
We wish to do immunocytochemistry on cultured myoblasts and myotubes;
all the antibodies that we wish to examine do not withstand fixation,
and we are assuming that we need to approach this from a cryo point of
view. There have been a couple of reports on their localization using
freeze-fracturing (and I know even less about that).
Before we start getting info and quotes from manufacturers, I'd like
some more information about this and other possible applications.
Would someone be able to give me a ball-park figure of how much it would
cost to perform cryoultramicrotomy taking into account technical time,
and consumables (minus the cost of primary antibody)?
Thanks in advance for any advice
Ronnie Houston
Texas Scottish Rite Hospital for Children
Dallas, TX 75219





From: Luc Harmsen :      anaspec-at-icon.co.za
Date: Tue, 15 Dec 1998 08:13:10 +0200
Subject: ISI SR50

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Yes no problem.
That is of course if the gun was an oil filled type. If it is the older =
epoxy type you will have to first get rid of the epoxy and clean it up =
before using some other HV epoxy to fill it again.
Luc Harmsen=09
Anaspec, South Africa
International technical support on microscopy.
Tel: +27 (0) 11 476 3455
Fax:+27 (0) 11 476 7290
anaspec-at-icon.co.za


-----Original Message-----
} From: "Corvos-at-aol.com"-at-sparc5.microscopy.com =
[SMTP:"Corvos-at-aol.com"-at-sparc5.microscopy.com]
Sent: Tuesday, December 15, 1998 3:18 AM
To: Microscopy-at-sparc5.microscopy.com


LUC,

Can you confirm again that Santovac 5 is OK to use on the gun... =20

Regards,

Walter Protheroe






From: Jim J Darley :      jim-at-proscitech.com.au
Date: Tue, 15 Dec 1998 21:17:34 +1100
Subject: blood stain; DNA vs cytoplasm

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I received an inquiry asking what would be a most suitable
stain
for student's to perform a quick blood stain to
differentiate
lymphocytes, macrophages, neutrophils etc (basically a DNA
vs cytoplasm
stain).
My background is mostly in EM and so I have trouble
thinking past Toluidine Blue O, differentiated with acid
alcohol. That would do, but I expect there is a better
alternative for smears.
Jim Darley

ProSciTech Microscopy
PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Phone +61 7 4774 0370 Fax: +61 7 4789 2313
Great microscopy catalogue, 500 Links, MSDS, User Notes
********************** www.proscitech.com.au *****








From: Sergey V. Lavristchev :      lavr-at-lst.gpi.ru
Date: Tue, 15 Dec 1998 15:10:08 +0300
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 15 Dec 98 08:41:14 -0500
Subject: FEG,ESEM or LV-SEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Biological SEM users:
I got very little response to the message I posted a week ago (see
below). It appears that at the present time there is not a great deal of
biological data from FEG-SEMs. To help us determine the best instrument to
consider for biological data accumulation, I would also like to explore
the pros and cons of the FEI/Philips'ESEM with the low vacuum instruments
available from other manufacturers. I have the brochures, etc from a number
of companies so am not looking for more of the same. I also have looked
into the basic design differences and detector differences.

What I would appreciate is comments from users based on their
experiences.
Why did you choose the instrument you did?
What type of information are you getting from your instrument of choice
that you could not get from a conventional SEM?
Do you also have a cryo unit and x-ray detector and, if so, what are the
limitations for their use in low vacuum modes?
How many hours per week is your instrument used and what percentage is in
the low vacuum mode?
Do you have other conventional SEMs or is this your only instrument?

Any other information would be appreciated. If the response is adequate,
and off-line, I will try to summerize.

Thanks,
Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057

-----------------------------------------------
Earlier post (12/7/98):
I would like to hear from microscopists who have used FEG-SEM's and
particularily the semi-in-lens models for biological applications. I
would
also appreciate info on recent papers where this instrumentation was
critical to the biological research reported. Also any recommendations
as to
ideal samples for testing these microscopes for biological application?
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057






From: Rozeveld, Steve (SJ) :      SJROZEVELD-at-dow.com
Date: Tue, 15 Dec 1998 12:00:24 -0600
Subject: PtSi2 TEM standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
(InterLock SMTP Gateway 3.0 for Microscopy-at-sparc5.microscopy.com);
Tue, 15 Dec 1998 12:00:27 -0600
Received: by na2.dow.com (Internal Mail Agent-1);
Tue, 15 Dec 1998 12:00:27 -0600
Message-Id: {D209047D0257D21191770000F8C991D916BB15-at-mante35a.nam.dow.com}




Could anyone suggest a source for PtSi2 ? (powder { 200 micron is
preferred ). I only need a few grams.
I thin film (i.e. similar to the NiOx TEM standard) would be ideal. Yes
- I did check with Goodfellows and Aesar. They will
provide the material as a special order but the minimum order was higher
than I needed.
thanks, Steve Rozeveld

Steve Rozeveld
The Dow Chemical Company
Analytical Sciences Laboratory
1897 Bldg., Door E43
Midland, MI 48667
517-636-5167 office
517-638-6443 fax






From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Tue, 15 Dec 1998 12:03:39 -0600
Subject: Kevex Delta system available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have a Kevex Delta V EDS system available for parts. It is available to
anyone willing to pay for the shipping.

We have kept the detector, amplifier, and HV power supply but have replaced
the computer section with a new Windows-based system.

The system contains a PDP-11/73 processor with 2-4 MB of memory, 1 10 MB
bernoulli drive, a 170 MB SCSI hard drive, and an ethernet card with
software. It has the full complement of Kevex software including Quantex,
Advanced Imaging, and Automated Image Analysis. It has both RT-11 and TSX
operating software.

If you or someone you know is interested, contact me via one of the means
below.

----------------------------------------------------
Warren E. Straszheim
23 Town Engineering
Iowa State University
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
http://www.marl.iastate.edu

electron microscopy, x-ray analysis, image analysis, computer applications






From: David Knecht :      knecht-at-uconnvm.uconn.edu
Date: Tue, 15 Dec 1998 14:35:52 -0500
Subject: oil overlay to prevent evaporation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have had some success with overlaying oil on a culture chamber in order
to prevent evaporation when using a 37=B0 heating plate for open chamber
microscopy. So far, I have used only light mineral oil. The problem is
that you have to put quite alot of oil on top to cover the surface. I am
wondering if anyone has found an oil that spreads better to form a barrier
with less volume.

Dr. David Knecht
Department of Molecular and Cell Biology
University of Connecticut
75 N. Eagleville Rd.
U-125
Storrs, CT 06269
Knecht-at-uconnvm.uconn.edu
860-486-2200
860-486-4331 (fax)







From: oshel-at-terracom.net (Philip Oshel)
Date: Tue, 15 Dec 1998 15:16:50 -0600
Subject: Re: blood stain; DNA vs cytoplasm

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jim,

This question would be best asked on Histonet--you'd get lots of quick
replies from folks who do blood smears and any other histo work for living.
I've included the Histonet listserver information below. Including some who
teach histotech and write texts on it.

If you can find a copy of "Staining Procedures" (Biological Stain
Commission), there are several stains given. Any histo text ought to have
these as well. The usual stains last time I did this were Wright's and
Giemsa for thin film. If you receipes, let me know--I can send you these
and a few others. Mind, mine are older ones.

Phil

WHO RUNS HISTONET?
The list is run by Linda R. Margraf, M.D. and Herb K. Hagler, Ph.D. using
hardware and software owned by the University of Texas Southwestern Medical
School, Department of Pathology in Dallas, Texas. If you have any questions
or problems with Histonet please contact Linda Margraf at
LMargraf-at-childmed.dallas.tx.us.

HOW DOES THE LIST WORK?
This server, unlike many systems, uses ONLY ONE ADDRESS to send commands to
the computer and to post messages. The server will recognize commands sent
in the SUBJECT line of the message and only when they are spelled exactly
as listed below. Anything not identified as a command will be circulated to
EVERYONE on the list.

The following is a list of commands the server recognizes:

subscribe
Your address will be added to the list of subscribers. You will then be
able to send messages to this list that will be forwarded to all other list
subscribers. You will begin to receive all messages sent to the list by
other subscribers.

subscribe digest
Your address will be added to the list of subscribers who receive a digest
instead of each forwarded message. A digest is a compilation of all the
messages received in a 24 hour period. It is sent to the digest subscribers
every night after midnight. Digest subscribers can post and respond to
messages the same as "real-time" subscribers.

digests
A list of available digests will be returned to you. Histonet stores old
messages as daily digests for approximately three months. To read previous
messages, copy the list of available digests, mark the dates of interest
and return it to the server.

unsubscribe
Your address will be removed from the list of subscribers. You will no
longer be able to send messages to the members of the list.

help
A list of the commands recognized by the server will be returned to you.

} I received an inquiry asking what would be a most suitable
} stain
} for student's to perform a quick blood stain to
} differentiate
} lymphocytes, macrophages, neutrophils etc (basically a DNA
} vs cytoplasm
} stain).
} My background is mostly in EM and so I have trouble
} thinking past Toluidine Blue O, differentiated with acid
} alcohol. That would do, but I expect there is a better
} alternative for smears.
} Jim Darley
}
} ProSciTech Microscopy
} PLUS
} PO Box 111, Thuringowa QLD 4817 Australia
} Phone +61 7 4774 0370 Fax: +61 7 4789 2313
} Great microscopy catalogue, 500 Links, MSDS, User Notes
} ********************** www.proscitech.com.au *****

****be famous! send in a tech tip or question***
Philip Oshel
Technical Editor, Microscopy Today
PO Box 620068
Middleton, WI 53562
(608) 833-2885
oshel-at-terracom.net








From: gorry2+-at-pitt.edu () (by way of Nestor J. Zaluzec)
Date: Tue, 15 Dec 1998 15:19:19 -0600
Subject: Help on basics of fluorescent microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Email: gorry2+-at-pitt.edu
Name: Michael Gorry
School: University of Pittsburgh

Question: Can you recommend any books or journal articles about the basics
of fluorescent microscopy photography? I know nothing about cameras and am
certain that I could improve the quality of the pictures that I have been
taking if I knew more about the camera and exposure times, aperature
sizing, etc. It is a Nikon microphot-FXA.

I could find no introductory information from the Nikon people.

Any help would be appreciated.

Thanks

Mike

Michael Gorry
E1105 BST
University of Pittsburgh
Arthritis Institute
Pittsburgh, PA 15261

gorry2+-at-pitt.edu

---------------------------------------------------------------------------







From: RCHIOVETTI-at-aol.com
Date: Tue, 15 Dec 1998 23:25:48 EST
Subject: Re: Help on basics of fluorescent microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In a message dated 98-12-15 16:38:37 EST, gorry2+-at-pitt.edu writes:

{ { Question: Can you recommend any books or journal articles about the basics
of fluorescent microscopy photography? } }

Mike,

There is a small book that was produced by Wild Leitz in Germany, entitled
"Fluorescence Microscopy: Principles, Instruments, Applications," written by
Eberhard Becker. Tips and tricks of photography are discussed on a couple of
pages in the book.

Are you having troubles with overexposed red fluorescence? Depending on the
vintage of the scope and the photodetector, there may be problems with reduced
sensitivity of the detector in the red end of the spectrum. The result:
seriously overexposed and washed-out photos if you set the camera for the
correct film speed.

You can compensate for this by multiplying the film speed (ASA) by about 4 or
adding about 6 to the DIN number when you take the photographs. This and
other topics are addressed in the book.

There is a chance that Leica still has some of these books. Call Leica
Customer Service at 1-800-248-0123 to inquire.

Hope this helps!

Cheers,

Bob
****************************************
Robert (Bob) Chiovetti, Ph.D.
President
Microimaging Technologies, Inc.
Tucson, Arizona USA
Tel. / Fax (520) 546-4986
rchiovetti-at-aol.com
Manufacturers' Representatives
Systems Integrators
Analog & Digital Imaging Systems
Research Microscopy Products
*****************************************





From: Robert Santoianni :      Robert_Santoianni-at-emory.org
Date: Wed, 16 Dec 1998 07:02:26 -0500
Subject: IF Photography

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To Michael Gorry, U of Pitt: re: Fluorescence Photomicrography

Your FXA may have too many "bells and whistles" for you to plow
through to understand what's happening with your photography. With
our rather simple set up, Nikon Optiphot with Microflex UFX II, I get
excellent pictures using Kodak Professional Ektachrome 1600, shot at
1600. Make sure your camera can spot meter, that is to be able to read
the brightest part of your field and not average the entire field. This way,
you meter the very strongest signal, set that value in the memory of the
timer, reposition your slide for the most pleasing shot and then expose.
Background should be dark to black...if you have flares or glare, there is
probably a diaphram somewhere between the lamphouse and the
microscope to stop down (but not to the point of vignetting your picture).
Good luck! I have an FXA instruction manual with a photography section
if you need a copy.
Bob Santoianni
Emory University Hospital
Atlanta, GA
robert_santoianni-at-emory.org





From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 12/15/98 4:19 PM
Subject: RE: Help on basics of fluorescent microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


A great, inexpensive and very readable initial sourse is : Herman,
B.(1998) Fluoresence Microscopy.BIOS Scientific Publishers Limited. ISBN:
0-387-91551-6.
This is part of the Royal Microscopy Society series published by
Springer-Verlag (volume 40).
It can be purchased through the publishers. I have often been able to
get books from this series through the publication "Microscopy Today".
This book gives an extensive list of references about equipment and
methods in fluorescence microscopy for standard, confocal and multi-photon
microscopy.

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057

--------------------------------------



Email: gorry2+-at-pitt.edu
Name: Michael Gorry
School: University of Pittsburgh

Question: Can you recommend any books or journal articles about the
basics
of fluorescent microscopy photography? I know nothing about cameras and
am
certain that I could improve the quality of the pictures that I have been
taking if I knew more about the camera and exposure times, aperature
sizing, etc. It is a Nikon microphot-FXA.

I could find no introductory information from the Nikon people.

Any help would be appreciated.

Thanks

Mike

Michael Gorry
E1105 BST
University of Pittsburgh
Arthritis Institute
Pittsburgh, PA 15261

gorry2+-at-pitt.edu

--------------------------------------------------------------------------
-





RFC822 header
-----------------------------------

RECEIVED: from SF_Database by POP_Mailbox_-1298363885 ; 15 DEC 98
21:50:11 UT
Received: from SPARC5.MICROSCOPY.COM by mailcenter.btny.purdue.edu
with SMTP (QuickMail Pro Server for MacOS 1.0.3); 15 DEC 98
21:50:02 UT
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com
(8.6.11/8.6.11) id PAA01759 for dist-Microscopy; Tue, 15 Dec 1998 15:08:29 -0600
Received: from no_more_spam.com (Sparc5 [206.69.208.10]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id PAA01756 for
"MicroscopyFilteredEmail-at-msa.microscopy.com"; Tue, 15 Dec 1998 15:07:58 -0600
Received: from [206.69.208.21] (mac21.zaluzec.com [206.69.208.21]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id PAA01749 for
{Microscopy-at-MSA.Microscopy.Com} ; Tue, 15 Dec 1998 15:07:45 -0600
X-Sender: zaluzec-at-microscopy.com
Message-Id: {v03007803b29c82bcad0e-at-[206.69.208.21]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
Date: Tue, 15 Dec 1998 15:19:19 -0600
To: Microscopy-at-Sparc5.Microscopy.Com
From: gorry2+-at-pitt.edu () (by way of Nestor J. Zaluzec)
Subject: Help on basics of fluorescent microscopy
Errors-to: Microscopy-request-at-sparc5.microscopy.com








From: csedax-at-alpha.arcride.edu.ar
Date: Wed, 16 Dec 1998 11:05:08 -2359
Subject: OM and SEM: they don't agree! (Thanks about the soft cheese samples!)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all!

First of all, thanks a lot for all the answers to my message regarding
problems while coating soft cheese samples. I talked to the interested persons
about your suggestions, those who also prepare the sample, I do the coating.
Indeed, it seems the problem is the temperature raise while coating (a big
evaporator adappted for sputtering). On the other hand the sample needs
another treatment for drying.
Thanks again for spending part of your time writing back to the net!


NEW QUESTION:

This time my question is related to large differences on measuring the
width of fibers seen from the surface of different kinds of papers.
We have taken micrographs with the SEM of a grid for calibrating at 300X, the
error in the calibration was found about 30% (before any calibration
correction). At the same time we took micrographs for the different samples at
the same magnification, under identical instrument conditions.

On the other hand, the same samples (before and after Au coating) were
observed under an optical microscope which was just calibrated. The width of
the same fibers showed to be 4 times longer than with the SEM!!!!! Widths in
the range of 10 - 20 microns on the SEM measure between 30-80 microns on the
OM.

Did any of you happen to have such an uncertainty on measurements? I
would trust the OM, but I have no arguments for saying that the SEM is wrong.

I would very much appreciate any comments

Many thanks in advance again.



Silvia Montoro
Centro Regional de Investigaciones y Desarrollo
Guemes 3450
3000 Santa Fe
Argentina
csedax-at-arcride.edu.ar








From: B.Laube-at-biologie.uni-bielefeld.de
Date: Wed, 16 Dec 1998 15:10:43 GMT+0100
Subject: Chromatography plate sectioning for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all, has anybody experience cutting silicified chromatographic
plates for TEM ultrastructure studies? We are afraid of destroying
our diamond knife, is this a serious fear or does it not matter? The
silicified graines are 50-60µm laying on a polyethylen sheet.
Thanks, Bw

Bernward Laube
University of Bielefeld
Faculty of Biology
Department Plant Morphology and Cell Ultrastructure
Universit„tsstrasse 25
Germany 33615 Bielefeld
phone: 0521 1065592
fax: 0521 1066039
e-mail: b.laube-at-biologie.uni-bielefeld.de
http://www.uni-bielefeld.de/biologie/Pflanzenmorphologie/abt.1htm#ws





From: B.Laube-at-biologie.uni-bielefeld.de
Date: Wed, 16 Dec 1998 15:14:35 GMT+0100
Subject: TEM/in situ hybridization

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all, has anybody experience or works with in situ hybridization
techniques and DNA/RNA negative staining for TEM examinations? Thanks for
help , Bw
Bernward Laube
University of Bielefeld
Faculty of Biology
Department Plant Morphology and Cell Ultrastructure
Universit„tsstrasse 25
Germany 33615 Bielefeld
phone: 0521 1065592
fax: 0521 1066039
e-mail: b.laube-at-biologie.uni-bielefeld.de
http://www.uni-bielefeld.de/biologie/Pflanzenmorphologie/abt.1htm#ws





From: Krzysztof Jan Huebner :      hubner-at-czapla.IOd.krakow.pl
Date: Wed, 16 Dec 1998 16:15:06 +0100 (MET)
Subject: news about "Color 1999" - now "Color 2000"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




15 December 1998

Dear Sirs,

I wish to let you know that due to reasons of the technical and
organisational nature it has been decided to shift the Conference Color
1999 to the year 2000 and hold it under the heading of Color 2000. I hope
that this "Royal Year" will prove to be more opportune for us - the
organizers, and for you - the participants. It will enable us to better
prepare this event, while you will have more time to prepare your
interesting papers.
Due to the above, all the dates stated previously will be shifted by
approximately one year, too.
Soon we are going to send more detailed agenda.

Wishing you Happy Christmas and Prosperous New Year 1999

I remain

Krzysztof Jan Huebner

{hubner-at-IOd.krakow.pl} :-)

FOUNDRY RESEARCH INSTITUTE
Research Materials Department
Manager of Structural and Physical Research Laboratory
str. Zakopianska 73 Call (*48 12) 2665022 ext.356
30-418 KRAKOW - POLAND Fax (+48 12) 2660870






From: William McManus :      billemac-at-bioserver.vsb.usu.edu
Date: Tue, 15 Dec 1998 12:43:33 -0700
Subject: FEG,ESEM or LV-SEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Debby;

The decision on which microscope to purchase depends on what type of work
and samples your lab will be handling. If you will be doing a lot of high
resolution work on fixed samples a FEG is great. We have a field emission
Hitachi SEM, an S4000. For low voltage work on biological samples it makes
fantastic images, but they need to be fixed, dried and coated. They have
exquisite detail, but if your sample cannot withstand the preparation or
your client is unwilling to pay for all the preparation, the results will be
disappointing. I do not have much experience with ESEMs, but from what I
have seen, I am not very impressed. They are the ultimate in ease for
looking at unfixed, wet samples, but the images from the ones I have seen
have looked pretty blah (a technical term). I think this is not just a
function of the image formation system in the scope, but of what you are
actually seeing, water. With all of the water present in a liquid form, it
flows and covers up the surface detail. I think that the combination of a
FEG with cryostage would be excellent compromise. It would allow looking at
unfixed samples, with everything in place without the time and expense of
the preparation, with all of the advantages of the FEG. It would also keep
the water were it was at the time of freezing. With sublimation, excess
water can be removed, thus exposing surface structures which may have been
covered. If you plan on looking at pollen and spore distributions, a
cryostage can make all the difference.
Bill

William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920


-----Original Message-----
} From: Debby Sherman [mailto:sherman-at-btny.purdue.edu]
Sent: Tuesday, December 15, 1998 6:41 AM
To: message to: MSA list


Biological SEM users:
I got very little response to the message I posted a week ago (see
below). It appears that at the present time there is not a great deal of
biological data from FEG-SEMs. To help us determine the best instrument to
consider for biological data accumulation, I would also like to explore
the pros and cons of the FEI/Philips'ESEM with the low vacuum instruments
available from other manufacturers. I have the brochures, etc from a number
of companies so am not looking for more of the same. I also have looked
into the basic design differences and detector differences.

What I would appreciate is comments from users based on their
experiences.
Why did you choose the instrument you did?
What type of information are you getting from your instrument of choice
that you could not get from a conventional SEM?
Do you also have a cryo unit and x-ray detector and, if so, what are the
limitations for their use in low vacuum modes?
How many hours per week is your instrument used and what percentage is in
the low vacuum mode?
Do you have other conventional SEMs or is this your only instrument?

Any other information would be appreciated. If the response is adequate,
and off-line, I will try to summerize.

Thanks,
Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057

-----------------------------------------------
Earlier post (12/7/98):
I would like to hear from microscopists who have used FEG-SEM's and
particularily the semi-in-lens models for biological applications. I
would
also appreciate info on recent papers where this instrumentation was
critical to the biological research reported. Also any recommendations
as to
ideal samples for testing these microscopes for biological application?
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057






From: Laura.Rhoads-at-wku.edu (Laura Rhoads)
Date: Wed, 16 Dec 1998 11:33:49 -0600
Subject: Looking for Olympus BH2 parts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Everyone!

I'm looking for some parts for the epifluorescence system of an Olympus BH2
with the old-style cylinder for holding the filters. Institutional policy requires
that we have no equipment less than a century old so Olympus no longer stocks these
items. They include:

the little cylinder that looks like part of a submarine periscope, with the filters mounted
in a horizontal configuration. I need a spare because my colleagues need the current filters
in it (for blue and green excitation), and I would like to do some DAPI. The only number on the
cylinder was H92103.

the little "bubble-wand" looking device which is mounted with the filters in a vertical position
that slides between two positions for blue and green. Again, I'd like a spare to place my DAPI
related filters in.

If anybody has a set of these that they would like to give away, sell, barter, etc. or
they know where I might find these, please contact me.

Happy Holidays!

Laura


************************************************************
Yes, it's true. The inmates ARE running the asylum...
************************************************************
Laura Rhoads
Electron Microscopy Facility Director
Department of Biology
Western Kentucky University
1 Big Red Way
Bowling Green, KY 42101-3576

(502) 745-6501 (502) 745-6856 fax







From: Woody.N.White-at-mcdermott.com
Date: Wed, 16 Dec 1998 10:10:00 -0600
Subject: Listserver reception question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Is it my system / mail client or did anyone else receive several
messages lately with no readable text?

For example, the recent mail from Gary Easton (Older SEM needs a
good home!), was a plain brown field with no text displayed.

Woody





From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Thu, 17 Dec 1998 03:14:19 -0500
Subject: OM and SEM: they don't agree! (Thanks about the soft cheese samples!)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Interesting your findings but not a shock to me! SEM calibrations,
particularly at low magnification may be at least 25% out. When did you
last have the microscope serviced by an experienced technician?

Remember that light and electron imaging reactions do differ. I am not a=
n
expert on light but I believe I understand what happens with electrons. =
I
would consider the LM magnifications to be more accurate but read on for
why the SEM has problems.

A conventional SEM image at less than 1,000X contains a great deal of BSE=

which either go straight into the Everhart-Thornley detector or are
converted to SE by bouncing off components in the chamber. Remember this
means sub surface information.

Electron images will have been produced from a volume of material that
differs from that of the LM due to penetration through and into the
material. You talk of fibres, one assumes they are hairy and that the L=
M
sees the hairs whilst the SEM does not? If you use the SEM at } 10kV you
will certainly have a considerable amount of penetration through the hair=
s
and as a result you just will not see them. If you or a helper has the
skill try 5 or even 2kV when you will see the hairs.

In the SEM business one becomes familiar with seeing things in light that=

do not show up with electrons and of course the reverse, its just part of=

the fun of this super subject.

Good luck

Steve Chapman

Senior Consultant E.M.
Protrain, 16 Hedgerley, Chinnor, Oxford OX9 4TN, England.
Tel & Fax 44 (0)1844 353161
Web Site - http://ourworld.compuserve.com/homepages/protrain
For Consultancy and Courses in Electron Microscopy World Wide





From: duchesne-at-alphixs.mpip-mainz.mpg.de
Date: Thu, 17 Dec 1998 09:48:27 +0100
Subject: Chromatography plate sectioning for TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} From: "B.Laube-at-biologie.uni-bielefeld.de"-at-Sparc5.Microscopy.Com
Organization: Fak. Biologie, Uni Bielefeld
To: Microscopy-at-Sparc5.Microscopy.Com
Date sent: Wed, 16 Dec 1998 15:10:43 GMT+0100


Dear all, has anybody experience cutting silicified chromatographic
plates for TEM ultrastructure studies? We are afraid of destroying
our diamond knife, is this a serious fear or does it not matter? The
silicified graines are 50-60=B5m laying on a polyethylen sheet.
Thanks, Bw

Bernward Laube
University of Bielefeld
Faculty of Biology
Department Plant Morphology and Cell Ultrastructure
Universit=84tsstrasse 25
Germany 33615 Bielefeld
phone: 0521 1065592
fax: 0521 1066039
e-mail: b.laube-at-biologie.uni-bielefeld.de
http://www.uni-bielefeld.de/biologie/Pflanzenmorphologie/abt.1htm#ws

sectioning of any matter is gonna harm a diamond knife with time...
the major problem in this case seems to be that some grains will be
pulled out of the matrix or away from the substrate, which will
a: locally strain the blade normal to the sectioning direction (weak
point of every diamond knife) and
b: lead to mostly useless sections.
I would try to somehow fix the grains (e.g. with a hard embedding
material on top) and preferably use an outer edge of a 45 degrees knife at=

fast speed.
Alexander Du Chesne, M.S., Ph.D.
Max-Planck-Institut f=FCr Polymerforschung
PF 3148, 55021 Mainz
Tel. 0049 6131 379 195
Fax 0049 6131 379 100
E-mail: duchesne-at-mpip-mainz.mpg.de





From: Chris Gilpin :      cgilpin-at-fs1.sem.man.ac.uk
Date: Thu, 17 Dec 1998 10:46:46 -0000
Subject: FEG,ESEM or LV-SEM?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


1) Why must you make a choice between these different instruments?
The FEGESEM allows imaging in either ESEM mode or hivac - user choice.
ESEM has come a long way since its beginnings and is worthy of full
investigation.
There is no need to have surface water present when imaging in ESEM mode.
Control of the specimen chamber environment allows an experienced user to
remove surface water without dehydrating the sample.

2)The other factor which has not been mentioned when trying to relate OM to
SEM is the effect of vacuum on the sample. There has been some nice work on
textile fibres by David Taylor in Leeds using SEM and ESEM on the same
fibres showing what must be high vacuum effect.


Chris

Chris Gilpin
Experimental Officer
Biological Sciences EM Unit
G452 Stopford Building
Oxford Road
Manchester
M13 9PT
phone +44 0161 275 5170
Fax +44 0161 275 5171 Chris Gilpin
http://www.empgu.man.ac.uk


Debby;

The decision on which microscope to purchase depends on what type of work
and samples your lab will be handling. If you will be doing a lot of high
resolution work on fixed samples a FEG is great. We have a field emission
Hitachi SEM, an S4000. For low voltage work on biological samples it makes
fantastic images, but they need to be fixed, dried and coated. They have
exquisite detail, but if your sample cannot withstand the preparation or
your client is unwilling to pay for all the preparation, the results will be
disappointing. I do not have much experience with ESEMs, but from what I
have seen, I am not very impressed. They are the ultimate in ease for
looking at unfixed, wet samples, but the images from the ones I have seen
have looked pretty blah (a technical term). I think this is not just a
function of the image formation system in the scope, but of what you are
actually seeing, water. With all of the water present in a liquid form, it
flows and covers up the surface detail. I think that the combination of a
FEG with cryostage would be excellent compromise. It would allow looking at
unfixed samples, with everything in place without the time and expense of
the preparation, with all of the advantages of the FEG. It would also keep
the water were it was at the time of freezing. With sublimation, excess
water can be removed, thus exposing surface structures which may have been
covered. If you plan on looking at pollen and spore distributions, a
cryostage can make all the difference.
Bill

William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920


-----Original Message-----
} From: Debby Sherman [mailto:sherman-at-btny.purdue.edu]
Sent: Tuesday, December 15, 1998 6:41 AM
To: message to: MSA list


Biological SEM users:
I got very little response to the message I posted a week ago (see
below). It appears that at the present time there is not a great deal of
biological data from FEG-SEMs. To help us determine the best instrument to
consider for biological data accumulation, I would also like to explore
the pros and cons of the FEI/Philips'ESEM with the low vacuum instruments
available from other manufacturers. I have the brochures, etc from a number
of companies so am not looking for more of the same. I also have looked
into the basic design differences and detector differences.

What I would appreciate is comments from users based on their
experiences.
Why did you choose the instrument you did?
What type of information are you getting from your instrument of choice
that you could not get from a conventional SEM?
Do you also have a cryo unit and x-ray detector and, if so, what are the
limitations for their use in low vacuum modes?
How many hours per week is your instrument used and what percentage is in
the low vacuum mode?
Do you have other conventional SEMs or is this your only instrument?

Any other information would be appreciated. If the response is adequate,
and off-line, I will try to summerize.

Thanks,
Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057

-----------------------------------------------
Earlier post (12/7/98):
I would like to hear from microscopists who have used FEG-SEM's and
particularily the semi-in-lens models for biological applications. I
would
also appreciate info on recent papers where this instrumentation was
critical to the biological research reported. Also any recommendations
as to
ideal samples for testing these microscopes for biological application?
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057








From: jeremy.sanderson-at-pathology.oxford.ac.uk
Date: Thu, 17 Dec 1998 13:10:30 0000
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Nestor
Please unsubscribe me from the list

Best wishes to all

Jeremy.Sanderson-at-path.ox.ac.uk





From: Woody.N.White-at-mcdermott.com
Date: 12/16/98 3:43 PM
Subject: Re: Listserver reception question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thanks all, for the reply. Looks like I may have to take matters
into my own hands unless "the company" soon up-grades the ccMail v
0.001 :) we are using.

Woody


______________________________ Reply Separator
_________________________________


It must be your Email program
I can read it fine. It does have embedded HTML code
looks like he uses a WWW browser based program
to send/read his mail


Nestor

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: rgriffin-at-eng.uab.edu
Date: Thu, 17 Dec 1998 09:15:52 -0600
Subject: OM and SEM: they don't agree! (Thanks about the soft cheese

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
id {Y5R69H3S} ; Thu, 17 Dec 1998 09:15:54 -0600
Message-ID: {7EDF3EF1CA87D211AC160000C0A90FE5115B61-at-engem0.eng.uab.edu}
To: Microscopy-at-Sparc5.Microscopy.Com
samples!)


Also - when you calibrate and make measurements on the SEM (and TEM) you
need to make sure that you are at the eucentric position. Measurement
values will change substantially at other heights.

Robin

-----Original Message-----
} From: Steve Chapman [mailto:PROTRAIN-at-CompuServe.COM]
Sent: Thursday, December 17, 1998 2:14 AM
To: INTERNET:"csedax-at-alpha.arcride.edu.ar"-at-sparc5.microscopy.com;
American Soc List


Hi,

Interesting your findings but not a shock to me! SEM calibrations,
particularly at low magnification may be at least 25% out. When did you
last have the microscope serviced by an experienced technician?

Remember that light and electron imaging reactions do differ. I am not an
expert on light but I believe I understand what happens with electrons. I
would consider the LM magnifications to be more accurate but read on for
why the SEM has problems.

A conventional SEM image at less than 1,000X contains a great deal of BSE
which either go straight into the Everhart-Thornley detector or are
converted to SE by bouncing off components in the chamber. Remember this
means sub surface information.

Electron images will have been produced from a volume of material that
differs from that of the LM due to penetration through and into the
material. You talk of fibres, one assumes they are hairy and that the LM
sees the hairs whilst the SEM does not? If you use the SEM at } 10kV you
will certainly have a considerable amount of penetration through the hairs
and as a result you just will not see them. If you or a helper has the
skill try 5 or even 2kV when you will see the hairs.

In the SEM business one becomes familiar with seeing things in light that
do not show up with electrons and of course the reverse, its just part of
the fun of this super subject.

Good luck

Steve Chapman

Senior Consultant E.M.
Protrain, 16 Hedgerley, Chinnor, Oxford OX9 4TN, England.
Tel & Fax 44 (0)1844 353161
Web Site - http://ourworld.compuserve.com/homepages/protrain
For Consultancy and Courses in Electron Microscopy World Wide





From: wft03-at-health.state.ny.us
Date: Thu, 17 Dec 1998 10:33:30 -0500
Subject: Re: Listserver reception question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Woody,
We must have a better environment than you; our field was green :-).
We too
have received several textually-challenged messages.
Yours,
Bill Tivol







From: Bob Roberts :      Bob.Roberts-at-asu.edu
Date: Thu, 17 Dec 1998 09:42:02 -0700
Subject: Ergonomic EM Operators Chair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Recently there was a posting inquiring about a source/vendor for a
non-magnetic operators chair for EM's. We would be interested in locating a
vendor for this type of product.

Bob Roberts
Arizona State University
Center for Solid State Science
Tempe, Arizona 85287








From: Michael A. O'Keefe :      maokeefe-at-lbl.gov
Date: Thu, 17 Dec 1998 11:33:19 -0800
Subject: Re: Ergonomic EM Operators Chair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------9336CF1E225BC7D15F9DF850
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

Bob,
One positive response to the posting is from Lund.


Mike,
We are using the IKEA chair Procent. It works well with the GIF 100.
Jan-Olov Bovin


-Mike O'Keefe




Bob Roberts wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Recently there was a posting inquiring about a source/vendor for a
} non-magnetic operators chair for EM's. We would be interested in locating a
} vendor for this type of product.
}
} Bob Roberts
} Arizona State University
} Center for Solid State Science
} Tempe, Arizona 85287



--------------9336CF1E225BC7D15F9DF850
Content-Type: text/x-vcard; charset=us-ascii; name="vcard.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for O'Keefe, Michael
Content-Disposition: attachment; filename="vcard.vcf"

begin: vcard
fn: Michael O'Keefe
n: O'Keefe;Michael
org: {center} {img src=http://mercproj.lbl.gov/logo.gif} {br} {b} {big} National Center for Electron Microscopy {/big} {/b} {/center}
email;internet: maokeefe-at-lbl.gov
title: Deputy Head
x-mozilla-cpt: ;0
x-mozilla-html: TRUE
version: 2.1
end: vcard


--------------9336CF1E225BC7D15F9DF850--






From: Mark Blackford :      mgb-at-ansto.gov.au
Date: Fri, 18 Dec 1998 09:14:40 -0600
Subject: Re: Ergonomic EM Operators Chair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Bob Roberts wrote recently:

} Recently there was a posting inquiring about a source/vendor for a
} non-magnetic operators chair for EM's. We would be interested in locating a
} vendor for this type of product.

This is an issue that many labs are facing, including my own. If there is
a commercially available solution (especially in Australia) please let us
all know via this listserver.

If not then we are considering doing as an Australian colleague has, i.e.
buy a suitably ergonomic chair(s) and have our workshop replace magnetic
components with brass. Not a cheap solution but justifiable considering
our investment in a GIF.

Cheers,


Mark Blackford
TEM Group
Materials Division, ANSTO
PMB 1,
Menai, N.S.W.
Australia
2234

Phone 61 2 9717 3027
Fax 61 2 9543 7179

Disclaimer:
The views expressed in this E-mail message do not necessarily represent the
official views of ANSTO from which this message was conveyed.







From: Julie Neimark (BIO) :      jneimark-at-chuma.cas.usf.edu
Date: Thu, 17 Dec 1998 19:22:19 -0500 (EST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



unsubscribe






From: Rozeveld, Steve (SJ) :      SJROZEVELD-at-dow.com
Date: Thu, 17 Dec 1998 20:44:14 -0500
Subject: TEM standard

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




} ----------
} From: Rozeveld, Steve (SJ)
} Sent: Tuesday, December 15, 1998 1:00 PM
} To: 'Microscopy-at-MSA.Microscopy.Com'
} Subject: PtSi2 TEM standard
}
}
}
} Could anyone suggest a source for PtSi2 ? (powder { 200 micron is OK
} ). I only need a few grams.
} I thin film (i.e. similar to the NiOx TEM standard) would be ideal.
} Yes - I did check with Goodfellows and Aesar. They will
} provide the material as a special order but the minimum order was
} higher than I needed.
} thanks, Steve Rozeveld
}
} Steve Rozeveld
} The Dow Chemical Company
} Analytical Sciences Laboratory
} 1897 Bldg., Door E43
} Midland, MI 48667
} 517-636-5167 office
} 517-638-6443 fax
}
}





From: Michael A. O'Keefe :      maokeefe-at-lbl.gov
Date: Thu, 17 Dec 1998 17:56:21 -0800
Subject: Re: Ergonomic EM Operators Chair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------CFE8ABC3BAEDCEA362E887B1
Content-Type: text/plain; charset=us-ascii
Content-Transfer-Encoding: 7bit

Mark,
Jan-Olov Bovin from Lund writes:

We are using the IKEA chair Procent.
It works well with the GIF 100.

I checked on the www and found that this chair seems
to use no metal in its construction.
I find it is described as:
PROCENT swivel chair.
Adjustable backrest angle and
seat depth. Compression
moulded base for improved
comfort. Star base and frame of
reinforced fiberglass
polypropylene foam.
Lanna 100% wool in black,
blue, dark blue, dark green,
dark red, grey or yellow.
at:
http://www.ikea-usa.com/content/products/prods/PROCENT_CHAIR.asp

On the other hand, I have not tested it, nor even seen one.
-Mike


Mark Blackford wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Bob Roberts wrote recently:
}
} } Recently there was a posting inquiring about a source/vendor for a
} } non-magnetic operators chair for EM's. We would be interested in locating a
} } vendor for this type of product.
}
} This is an issue that many labs are facing, including my own. If there is
} a commercially available solution (especially in Australia) please let us
} all know via this listserver.
}
} If not then we are considering doing as an Australian colleague has, i.e.
} buy a suitably ergonomic chair(s) and have our workshop replace magnetic
} components with brass. Not a cheap solution but justifiable considering
} our investment in a GIF.
}
} Cheers,
}
} Mark Blackford
} TEM Group
} Materials Division, ANSTO
} PMB 1,
} Menai, N.S.W.
} Australia
} 2234
}
} Phone 61 2 9717 3027
} Fax 61 2 9543 7179
}
} Disclaimer:
} The views expressed in this E-mail message do not necessarily represent the
} official views of ANSTO from which this message was conveyed.



--------------CFE8ABC3BAEDCEA362E887B1
Content-Type: text/x-vcard; charset=us-ascii; name="vcard.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for O'Keefe, Michael
Content-Disposition: attachment; filename="vcard.vcf"

begin: vcard
fn: Michael O'Keefe
n: O'Keefe;Michael
org: {center} {img src=http://mercproj.lbl.gov/logo.gif} {br} {b} {big} National Center for Electron Microscopy {/big} {/b} {/center}
email;internet: maokeefe-at-lbl.gov
title: Deputy Head
x-mozilla-cpt: ;0
x-mozilla-html: TRUE
version: 2.1
end: vcard


--------------CFE8ABC3BAEDCEA362E887B1--






From: rice-at-mcc.com (Janet Rice)
Date: Thu, 17 Dec 1998 20:50:52 -0600
Subject: FESEM Retrofit?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



We currently have a Hitachi S570 with a LaB6 filament, however, my
management would like to see us get into semiconductor work. For this we
would need an FESEM. The questions was asked as to whether or not a
standard SEM could be retrofit as with the FE source. My best guess would
be "no", but perhaps someone out there has done this - or would know for
sure it can't be done. Any information would be welcome (and yes, it would
be nice to by a whole new instrument, but.....)

Janet Rice
MCC
Senior Member Technical Staff
rice-at-mcc.com
512-338-3266







From: Barbara Foster :      mme-at-map.com
Date: Fri, 18 Dec 1998 11:18:00 -0500
Subject: Re: Help on basics of fluorescent microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mike,


We have found two books very helpful, one by Kodak and one by Polaroid.


The one by Kodak is called "Photography Through the Microscope" and is by
John Delly. It should be available through most camera stores. We have a
few extras here, available for sale, if you can't find it locally.


The other is "Instant Photomicrography Through the Microscope" and should
be available through the Polaroid Tech Support hotline (I'm sorry - no
contact info - call 1-800-555-1212 for info; they are in Cambridge,
MA).


Hope this is helpful.

Barbara Foster

Consortium President

{color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
..Educating microscopists for greater productivity.


{/color} 125 Paridon Street Suite 102 Springfield, MA 01118

PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

Visit our web site { {http://www.MME-Microscopy.com/education}

******************************************************

{bigger} {bigger} MME {/bigger} {/bigger} is America's first national
consortium dedicated to

customized on-site training in all areas of

microscopy, sample preparation, and image analysis {bigger} .

{/bigger}




At 11:25 PM 12/15/98 EST, RCHIOVETTI-at-aol.com"-at-Sparc5.Microscopy.Com
wrote:

} ------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

} -----------------------------------------------------------------------.

}

}

} In a message dated 98-12-15 16:38:37 EST, gorry2+-at-pitt.edu writes:

}

} { { { { Question: Can you recommend any books or journal articles about the
basics

} of fluorescent microscopy photography? } }

}

} Mike,

}

} There is a small book that was produced by Wild Leitz in Germany,
entitled

} "Fluorescence Microscopy: Principles, Instruments, Applications,"
written by

} Eberhard Becker. Tips and tricks of photography are discussed on a
couple of

} pages in the book.

}

} Are you having troubles with overexposed red fluorescence? Depending on
the

} vintage of the scope and the photodetector, there may be problems with
reduced

} sensitivity of the detector in the red end of the spectrum. The
result:

} seriously overexposed and washed-out photos if you set the camera for
the

} correct film speed.

}

} You can compensate for this by multiplying the film speed (ASA) by about
4 or

} adding about 6 to the DIN number when you take the photographs. This
and

} other topics are addressed in the book.

}

} There is a chance that Leica still has some of these books. Call
Leica

} Customer Service at 1-800-248-0123 to inquire.

}

} Hope this helps!

}

} Cheers,

}

} Bob

} ****************************************

} Robert (Bob) Chiovetti, Ph.D.

} President

} Microimaging Technologies, Inc.

} Tucson, Arizona USA

} Tel. / Fax (520) 546-4986

} rchiovetti-at-aol.com

} Manufacturers' Representatives

} Systems Integrators

} Analog & Digital Imaging Systems

} Research Microscopy Products

} *****************************************

}

}

}







From: shAf :      mshaf-at-darkwing.uoregon.edu
Date: Fri, 18 Dec 1998 08:08:18 -0800
Subject: RE: FESEM Retrofit?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Janet asks ...

} ...
}
}
}
} We currently have a Hitachi S570 with a LaB6 filament, however, my
} management would like to see us get into semiconductor work.
} For this we would need an FESEM.
} The questions was asked as to whether or not a
} standard SEM could be retrofit as with the FE source. ...

The answer wouldn't be absolutely not ... after all, FE guns
are adapted to similar electron columns. I can think of 3 issues
which make the conversion quite expensive (... maybe other can
include others ...)

(1) although your LaB6 gun is ion pumped, FE will demand a better
ion pump and possibly more efficent 1st stage pumping.

(2) FE guns are said, maybe debatably, to be inherently unstable
.. regardless, FE instability is generally compensated for
brightness/contrast by an additional beam regulation component.

(3) The resulting increased resolution (... presumably this is why
you want FE ...) demands mu-metal electro-magnetic shielding.

... hope this helps :o)

cheerios, shAf

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - ICQ 210524
Geological Science's Electron Probe Facility - University of Oregon
mshaf-at-darkwing.uoregon.edu - http://darkwing.uoregon.edu/~mshaf/







From: Owen P. Mills :      opmills-at-mtu.edu
Date: Fri, 18 Dec 1998 11:39:51 -0500
Subject: 4pi SEII to Link eXL

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi:

I'd like to connect with someone who has tried, successfully (or not),
connecting 4pi's SEII card to a Link eXL XP2 pulse processor (PC1934).
Please contact me off line at the address below. TIA.

Owen



=============================
Owen P. Mills
Michigan Technological University
Metallurgical & Materials Engineering
Rm 512 MME Building
Houghton, MI 49931
906-487-2002
906-487-2934 FAX
opmills-at-mtu.edu





From: psic-at-uclink4.berkeley.edu (Paula Sicurello)
Date: Fri, 18 Dec 1998 09:32:00 -0800 (PST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





Please,




unsubscribe psic-at-uclink4.berkeley.edu







Happy Holidays everybody! :-)

Paula Sicurello
UC Berkeley
Electron Microscope Lab
psic-at-uclink4.berkeley.edu
phone: 510-642-2085
fax: 510-643-6207
http://biology.berkeley.edu/EML







From: Fagerland,Jane :      jane.a.fagerland-at-abbott.com
Date: Fri, 18 Dec 1998 11:45:27 -0600
Subject: Cryo TEM service needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Cryo TEM Service Work. I need a lab that is doing outside
contract cryo TEM or cryo TEM replica work. I have some samples th=
at I
need prepared and photographed.

Contact: Joanne M. Crudele
Unilever HPC USA
Joanne.Crudele-at-Unilever.com
or 847-734-3712
=





From: Yvan Lindekens :      yvan.lindekens-at-skynet.be
Date: Sat, 19 Dec 1998 11:42:02 +0100
Subject: Re: blood stain; DNA vs cytoplasm

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A simple Pappenheim-stain (May-Grunwald/Giemsa) is probably the simplest
way to make a "quick blood stain to differentiate...".

This stain is easy to perform, fast (about 10 - 15 min) and gives very good
differentiation of all blood-components!

Protocols can be found in any textbook on basic histological technique. If
you can't find them, let me know...

Yvan Lindekens.
----------
} Van: Jim J Darley {jim-at-proscitech.com.au}
} Aan: Microscopy-at-sparc5.microscopy.com {Microscopy-at-Sparc5.Microscopy.Com}
} Onderwerp: blood stain; DNA vs cytoplasm
} Datum: dinsdag 15 december 1998 11:17
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------
} I received an inquiry asking what would be a most suitable
} stain
} for student's to perform a quick blood stain to
} differentiate
} lymphocytes, macrophages, neutrophils etc (basically a DNA
} vs cytoplasm
} stain).





From: Yvan Lindekens :      yvan.lindekens-at-skynet.be
Date: Sat, 19 Dec 1998 11:31:38 +0100
Subject: Re: Help on basics of fluorescent microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


In addition to RCHIOVETTI-at-aol.com message: another interesting brochure on
fluorescent microscopy was made by Reichert in Vienna (now also Leica):

"Fluorescence microscopy with fluorochromes - recipes and tables"

A good chapter on fluorescent microscopy photography can be found in:

C. Van Duyn Jr: "Mikrofotografie", Focus Elsevier, The Netherlands (no
ISBN). This is in Dutch.

Some intresting notes on the subject can also be found in:

Goke, G: "Modene Methoden der Licht-mikroskopie", Kosmos Wissensschaft,
Franckh, Stuttgart 1988, ISBN: 3-440-05765-8. This is in German.

Hope this helps...

Yvan Lindekens.


----------
} Van: RCHIOVETTI-at-aol.com-at-Sparc5.Microscopy.Com
} Aan: gorry2+-at-pitt.edu; Microscopy-at-Sparc5.Microscopy.Com
} Onderwerp: Re: Help on basics of fluorescent microscopy
} Datum: woensdag 16 december 1998 5:25
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} In a message dated 98-12-15 16:38:37 EST, gorry2+-at-pitt.edu writes:
}
} { { Question: Can you recommend any books or journal articles about the
basics
} of fluorescent microscopy photography? } }
}
} Mike,
}
} There is a small book that was produced by Wild Leitz in Germany,
entitled...






From: Julian Smith III :      smithj-at-Winthrop.edu
Date: Sat, 19 Dec 1998 09:21:10 -0500
Subject: WTB:Used Nikon CF LWD 40x; Nikon fluo cubes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, all
We're looking for an LWD 40x phase objective and some filter cubes to go
with a used Nikon Diaphot with epi-fluorescence. If you have one or more
of these items, please contact me at the e-mail address below; please do
not reply to the list.

The subject line says nearly all of it. We're looking for a 40x
long-working distance phase objective with correction collar for a Nikon
diaphot:
Any one of the following three objectives will do:
} CF Achromat LWD DL 40x C (78816)
} CF N Plan Achromat ELWD DM 40xC (85053)
} CF Fluor LWD DM 40x C (85008)

We are also looking for blue, green and UV cubes for an epifluorescence
illuminator; standard Nikon will do, as these are going in an Opti-Quip slider

TIA
Julian Smith III
Biology
Winthrop University
Rock Hill SC
803-323-2246 (fax)
803-323-2111 x6427 (vox)
smithj-at-winthrop.edu

Julian P.S. Smith III
Biology
Winthrop University
Rock Hill, SC 29733
803-323-2111 x6427 (vox)
803-323-2246 (fax)







From: Gillmeister, Russ :      RGillmeister-at-sdms.usa.xerox.com
Date: Mon, 21 Dec 1998 08:22:54 -0500
Subject: Re: Ergonomic EM Operators Chair

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-----Original Message-----
} From: Doug Danielson [mailto:wddanie-at-ibm.net]
Sent: Saturday, December 19, 1998 1:48 PM
To: Microscopy-at-MSA.Microscopy.Com


Hi Mark, I must be getting real old as I remember my gransfather used to
make chairs out of plant material. Large plants I believe. They wood cut the
larger stems into structural members and glue the parts together into many
different shapes. I sure these were not as functional or beautiful as the
bent steel and plastic used today. Maybe you could find one of these
antiques for your purpose.
Russ

-----Original Message-----
} From: Mark Blackford [mailto:mgb-at-ansto.gov.au]
Sent: Friday, December 18, 1998 10:15 AM
To: Microscopy-at-sparc5.microscopy.com


Bob Roberts wrote recently:

} Recently there was a posting inquiring about a source/vendor for a
} non-magnetic operators chair for EM's. We would be interested in locating a
} vendor for this type of product.

This is an issue that many labs are facing, including my own. If there is
a commercially available solution (especially in Australia) please let us
all know via this listserver.

If not then we are considering doing as an Australian colleague has, i.e.
buy a suitably ergonomic chair(s) and have our workshop replace magnetic
components with brass. Not a cheap solution but justifiable considering
our investment in a GIF.

Cheers,


Mark Blackford
TEM Group
Materials Division, ANSTO
PMB 1,
Menai, N.S.W.
Australia
2234

Phone 61 2 9717 3027
Fax 61 2 9543 7179

Disclaimer:
The views expressed in this E-mail message do not necessarily represent the
official views of ANSTO from which this message was conveyed.







From: Dr. Dr. Ingrid Voigt-Martin :      voigtmar-at-mail.uni-mainz.de
Date: Mon, 21 Dec 1998 15:28:50 +0100 (MET)
Subject: Postdoc position available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} From January 1999 there will be a 3 year post-doc position available in the
electron microscopy group (Department of Physical Chemistry) at the
University of Mainz, Germany.
At present the group has 2 electron microscopes and a well-equipped specimen
preparation laboratory. In the summer of 1999 a new 300 kV Philips Tecnai 30
Electron Microscope with FEG and GIF will be installed.
Applications from physicists and chemists with Ph.D experience in electron
microscopy and a sound background in electron energy loss spectroscopy are
welcome.

Please send applications to

Dr. I.G. Voigt-Martin
Inst. f. Physikalische Chemie
Jacob Welder Weg 11
55099 Mainz






From: MICHAEL DELANNOY :      delannoy-at-welchlink.welch.jhu.edu
Date: Mon, 21 Dec 1998 11:20:16 -0500 (EST)
Subject: Equipment sale

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hello,
I have one Sorvall MT 5000 ultramicrotome that is in operating
condition, and a Durst Laborator (185) enlarger with all the lens' still
operational, for sale. If ineterested, E-mail or call with an offer.

Michael Delannoy
(410) 955 -1365

delannoy -at-welchlink.welch.jhu.edu






From: Ritchey, M Barry :      mbritch-at-sandia.gov
Date: Mon, 21 Dec 1998 10:50:58 -0700
Subject: Wanted: Polaroid Camera Controller for Leco-300

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



The Matic-Mot2 camera controller for my Leco 300 has died - it gives a
"Camera?" message and doesn't function. Anybody out there have this happen?
Any known source to have it fixed? The Leco folks say that parts are very
rare and the cost to fix would be $2-4K (USD).

If anyone has a spare Matic-Mot2 camera controller or knows of a fix, let me
know.

I decided to hang a Polaroid DMC on the microscope, but would still like to
have the capability to take 4X5 Polaroids.

Happy Holidays,

Barry Ritchey
mbritch-at-sandia.gov
Sandia Nat'l Labs - NM






From: DChernoff-at-aol.com
Date: Tue, 22 Dec 1998 00:04:41 EST
Subject: wet polishing wheel

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm looking for a used, inexpensive wet polishing wheel.
Contact Don Chernoff, (703) 849-1492, dchernoff-at-aol.com





From: Debashis Mukherji :      d.mukherji-at-tu-bs.de
Date: Tue, 22 Dec 1998 13:11:21 +0100 (MET)
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


******************************************
Dr-Ing. D. Mukherji
Institut fuer Werkstoffe
Technische Universitaet Braunschweig
Langer Kamp 8
38106 Braunschweig, Germany
Tel: (0531) 391 3063
Fax: (0531) 391 3058
******************************************






From: Steven W. Miller :      Steven_W_Miller-at-CompuServe.COM
Date: Tue, 22 Dec 1998 12:16:26 -0500
Subject: Microtomy of Ag Flakes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We would suggest that you take several approaches to evaluating your
suspect contamination.

Sectioning a few microns thick (this will be possible if the epoxy is no=
t
too hard) and looking for an organic that should not be there.

Consider that it may perish in a vacuum; try AFM imaging of the microtome=
d
surface; either LFM or phase imaging should do.

TEM/EDX on the thin sections may show you an inorganic layer that is too=

thin to give a usable signal in SEM mode. =


Let us know if we can help, we have a full applications laboratory for
preparing your samples in Tucson,AZ.

Regards,
Steve Miller
Director of Sales, North America

We are a commercial manufacturer of microtomes and other microscopy sampl=
e
preparation equipment.

Ventana/RMC
3450 S. Broadmont
Tucson, AZ 85713
520-903-9366 Phone
520-903-0132 Fax
RMC-at-RMC-Scientific.com

Website: RMC-Scientific.com/microtomes/





From: South Bay Technology :      Henriks-at-CompuServe.COM
Date: Tue, 22 Dec 1998 12:47:45 -0500
Subject: wet polishing wheel

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Don:

While we don't have a used polishing wheel, we do offer an inexpensive 8"=

wet polishing wheel (Model 900) which is less than $1,000. For informati=
on
you can see our web site at www.southbaytech.com or give me a call at
800-728-2233.

Best regards-

David =

Writing at 9:10:17 AM on 12/22/98
=

*************************************************************************=
**
************************

David Henriks TEL: =

800-728-2233 (toll free in the USA)
South Bay Technology, Inc. +1-949-492-2600
1120 Via Callejon FAX: +1-949-492-1499=

San Clemente, CA 92673 USA e-mail: henriks-at-southbaytech.com=


*************************************************************************=
**
************************

} } } } } Please visit us at http://www.southbaytech.com { { { { {

Manufacturers of precision sample preparation equipment and supplies for
metallography, crystallography and electron microscopy.


Message text written by INTERNET:"DChernoff-at-aol.com"-at-sparc5.microscopy.co=
m
} ------------------------------------------------------------------------=

The Microscopy ListServer -- Sponsor: The Microscopy Society of America =



I'm looking for a used, inexpensive wet polishing wheel.
Contact Don Chernoff, (703) 849-1492, dchernoff-at-aol.com

{






From: Ricky L Vaughn :      RLVAUGHN-at-MAIL.UNMC.EDU
Date: Tue, 22 Dec 1998 15:54:27 -0600
Subject: TEM Processers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello All

A while back someone asked about automatic tissue processers. I don't
recall seeing much responce. I have been asked to investigate pros and
cons for our pathology dept. TEM lab. I have literature on RMC and
Leica. We looked at one ten years ago and our only concern at that time
was possible drying artifact from tissue being lifted out of the propylene
oxide, and that there were not many people using them. Thanks for any
help.

Rick Vaughn
Univ. Nebr. Med. Ctr.
RLVAUGHN-at-MAIL.UNMC.EDU





From: Krzysztof Jan Huebner :      hubner-at-czapla.IOd.krakow.pl
Date: Wed, 23 Dec 1998 10:09:01 +0100 (MET)
Subject: zyczenia swiateczne

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Kochani,


Wam wszystkim z ktorymi wspolpracuje z okazji Swiat Bozego Narodzenia i
Nowego Roku 1999 skladam zycenia Duzo Zdrowia i Wszystkiego
Najlepszego dla Was i Waszych Najblizszych.

Niech kazdemu spelni sie przynajmniej jedno z tych cichych marzen ktore,
gdy wieczorem zasiada sie w fotelu to wraca - a gdy tak ....

Wszystkiego Najlepszego


Krzysztof Jan Huebner

{hubner-at-IOd.krakow.pl} :-)

FOUNDRY RESEARCH INSTITUTE
Research Materials Department
Manager of Structural and Physical Research Laboratory
str. Zakopianska 73 Call (*48 12) 2665022 ext.356
30-418 KRAKOW - POLAND Fax (+48 12) 2660870






From: Dmitri V. Sokolov :      sokolov-at-ryouko.rciqe.hokudai.ac.jp
Date: Wed, 23 Dec 1998 19:48:36 +0900
Subject: JEM-2010X manual in English

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi!

I am going to work with TEM JEM-2010X, but to start is hard for me because
of manuals, that are written in Japanese.
I need to make a chemical composition analysis of about 3x3 micron area on
semiconductor surface. And the only installation for such an analysis with
appropriate spatial resolution is JEM-2010X. To simplify the procedure I'd
like to avoid the common routine of sample preparations for TEM. But for
that I need to know if the mentioned installation has such an option to
observe the surface in non-transmission way.

Any suggestions on where to get the manual in English (printed, or in any
file format) would be highly appreciated.

Dmitri.

__________________________________________
Dmitri V. Sokolov
Doctor Course student
Research Center for Interface Quantum Electronics,
Hokkaido University, North 13, West 8, Kitaku,
Sapporo 060, Hokkaido, Japan

Phone 81-11-706-7174
Fax 81-11-716-6004
Home phone 81-11-261-2185

http://www.geocities.com/SiliconValley/Campus/1314
mailto:sokolov-at-ryouko.rciqe.hokudai.ac.jp - daily
mailto:dv_sokolov-at-hotmail.com - weekly
AOL Instant Messenger FalconDot
ICQ 9418072
__________________________________________







From: Krzysztof Jan Huebner :      hubner-at-czapla.IOd.krakow.pl
Date: Wed, 23 Dec 1998 15:39:37 +0100 (MET)
Subject: Re: merry christmates and "zyczenia swiateczne"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Merry Christmas and all the best for 1999.

and

If you have any wishes I wishe you to have them made true !!!



Krzysztof Jan Huebner

{hubner-at-IOd.krakow.pl} :-)

FOUNDRY RESEARCH INSTITUTE
Research Materials Department
Manager of Structural and Physical Research Laboratory
str. Zakopianska 73 Call (*48 12) 2665022 ext.356
30-418 KRAKOW - POLAND Fax (+48 12) 2660870

On Wed, 23 Dec 1998 VANDERHAEGHEN_JACQUES/CENTER_RD-at-bekaert.com wrote:

} Please if you send a e-mail do it in Englisch.
} But, I want to send you the best wisches for Christmas and a very happy
} new year.
}
} Jacques Vander Haeghen
} Belgium
} ----------
}
} Kochani,
}
}
} Wam wszystkim z ktorymi wspolpracuje z okazji Swiat Bozego Narodzenia i
} Nowego Roku 1999 skladam zycenia Duzo Zdrowia i Wszystkiego
} Najlepszego dla Was i Waszych Najblizszych.
}
} Niech kazdemu spelni sie przynajmniej jedno z tych cichych marzen ktore,
} gdy wieczorem zasiada sie w fotelu to wraca - a gdy tak ....
}
} Wszystkiego Najlepszego
}
}
} Krzysztof Jan Huebner
}
} {hubner-at-IOd.krakow.pl} :-)
}
} FOUNDRY RESEARCH INSTITUTE
} Research Materials Department
} Manager of Structural and Physical Research Laboratory
} str. Zakopianska 73 Call (*48 12) 2665022 ext.356
} 30-418 KRAKOW - POLAND Fax (+48 12) 2660870
}
}
}





From: mcarter :      mcarter-at-fccjvm.fccj.cc.fl.us
Date: Wed, 23 Dec 1998 10:43:16 -0500
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html









From: Ed Calomeni :      ecalomeni-at-mco.edu
Date: Wed, 23 Dec 1998 12:30:36 -0500
Subject: Re: TEM Processers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Ricky,

I have been using a RMC processor for 3 - 4 years now. It is used on a =
daily basis, yes a few problems and yes a few lost specimens. If =
possible do not process all of biospy received. I like to RMC because of =
the individual caps to reagents. Have not had any problems do to =
evaporation of solvents. =20

Best of Luck,

=20

Ed Calomeni
Dept. Pathology
Medical College of Ohio
3000 Arlington Avenue
Toledo, Ohio 43614

phone: 419-383-3484
fax: 419-383-3066
ecalomeni-at-mco.edu

Ed Calomeni
Dept. Pathology
Medical College of Ohio
3000 Arlington Avenue
Toledo, Ohio 43614

phone: 419-383-3484
fax: 419-383-3066
ecalomeni-at-mco.edu






From: James Pawley :      jbpawley-at-facstaff.wisc.edu
Date: Wed, 23 Dec 1998 21:06:14 -0500
Subject: Fourth course on 3D Microscopy of Living Cells next June: First

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

Just to start the New Year off on an up beat: I would like to announce:

The Fourth International Short Course on 3D Microscopy of Living Cells
June 16 - 27, 1999

and

The Post-course Workshop on 3D Image Processing
June 29- July 1, 1999

in association with the
BioSciences Microscopy Facility and Department of Computer Science

University of British Columbia, Vancouver, BC, Canada

Organized by Prof. James Pawley: University of Wisconsin-Madison


We expect to have at least 11, 3D microscope workstations, including
several using multi-photon excitation, for student use and there will be an
international faculty of 14. The schedule is basically
lecture/demonstration in the mornings, labs in the afternoons and personal,
live cell projects at night. Many contributors to this list are "alumnae."


Additional information is available from:

http://www.cs.ubc.ca/spider/ladic/course/bulletin.html

And if you would like to see what we did last year, try

http://www.cs.ubc.ca/spider/ladic/course/Course98/bulletin.html



APPLICATIONS
Applicants will complete a questionnaire to assess knowledge level,
field of interest and proposed personal, live-cell, project. Enrollment
will be limited to about 24 participants. Selection will be made on the
basis of background and perceived need. Those without previous LM
experience will be provided with basic texts to read before the course
begins. Application forms requesting information on field of interest and
level of experience may be down-loaded from the WWW site at


http://www.cs.ubc.ca/spider/ladic/course/forms/register.txt or obtained from:

Prof. James Pawley, Rm. 1235,
1117 W. Johnson Ave., Madison, WI, USA 53706.
Phone: 1-608-263-3147, Fax 1-608-265-5315,
Email: jbpawley-at-facstaff.wisc.edu


Application deadlines:

Application forms must be received for screening by March 1, 1999.
Successful applicants will be notified by April 1, and a deposit of 50%
must be received by April 15, 1999. In general, refunds of the deposit will
not be possible. The remainder is due before Registration.

DATES:
Applications must be received by Mar. 1/99
50% deposit due Apr. 15/99
Registration 4:00 - 7:00 pm Wednesday, June 16, 1999
Last class will end with lunch Sunday, June 27, 1999

3D Image Processing Workshop June 29- July 1, 1999

3D Course tuition (includes lunches): $1950 (US)
Workshop Tuition (includes lunches): $800 (US)

Room / board about $40/day

Hope that you all have a great holiday season!!

Cheers,

Jim Pawley





From: dmrelion-at-world.std.com (donald j marshall)
Date: Thu, 24 Dec 1998 08:57:16 -0500
Subject: penetration depth

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I realize this is not the best time to post a question to the list, with so
many people unsubscribed for the holidays, but here goes. The question has
to do with the generation of Cathodoluminescence (CL) in specimens as a
function of incident electron energy. Leverenz (1968) has a number,
apparently a calculated range, of 0.2 microns for a 6 keV electron in
Zn2SiO4 and says that the electron range is proportional to the square of
the electron energy in this range. Is there any good more recent data on the
penetration depth of electrons in minerals, e.g., silicates, carbonates, as
a function of the electron enery,in the range of 5kV to 100 kV?

Don Marshall

Donald J. Marshall
Relion Industries
P.O. Box 12
Bedford, MA 01730
Ph: 781-275-4695
FAX: 781-271-0252
email dmrelion-at-world.std.com

"A weed is a flower out of place."






From: MICHAEL DELANNOY :      delannoy-at-welchlink.welch.jhu.edu
Date: Thu, 24 Dec 1998 09:59:03 -0500 (EST)
Subject: Microtome service contracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Happy Holidays,
Does anyone know of a company or small contractor who will
service Reichert microtomes (other than the more expensive bigger
companies like Leica). I recall a company called Tek-net in N.J.
but am unsure. Salutaions,

Mike D







From: Sara Miller :      saram-at-acpub.duke.edu
Date: Thu, 24 Dec 1998 16:24:00 -0500 (EST)
Subject: Re: Microtome service contracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


TekNet is great. We've been very satisfied.

No commercial interest.


Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735






From: Dr P. Echlin :      pe13-at-cus.cam.ac.uk
Date: Fri, 25 Dec 1998 11:24:18 +0000 (BST)
Subject: Re: Microtome service contracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Dear Michael:

I know it's Christmas day, but I have opened all my prezzies and found
your message about penetration depth. You can calculate the average
depth of electron peneration using the Bethe range equation. See
equation 3.13 Page 88 of SEM-XRMA by
Goldstein,Newbury,Echlin,Joy,Romig,Lyman,Fiori & Lifshin. lenum Press NY
1992 which gives all the good stuff. Also use Electron Flight simulator
a piece of software which does all the sums for you.

Happy Christmas.

I'm back to the claret and Stilton

Patrick Echlin
Multi-Imaging Centre
Cambridge University
United Kingdom

On Thu, 24 Dec 1998,
MICHAEL DELANNOY wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Happy Holidays,
} Does anyone know of a company or small contractor who will
} service Reichert microtomes (other than the more expensive bigger
} companies like Leica). I recall a company called Tek-net in N.J.
} but am unsure. Salutaions,
}
} Mike D
}
}
}
}






From: Sally Stowe :      stowe-at-rsbs.anu.edu.au
Date: Sun, 27 Dec 1998 17:36:29 +1100
Subject: Wanted - Cold stage for Philips 430 TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Happy Christmas everybody
Does anybody have an unwanted liquid nitrogen-cooled cold stage suitable =
for a Philips 430 TEM, with retractable protective blades? We have =
temperature control and transfer systems for Oxford and Gatan systems on =
other TEMS, so may only need the specimen holder itself. But we are =
looking for the type of holder that can be used to transfer a cold =
specimen into the microscope, rather than just cool it down within the =
column.=20
Yours in hope
Sally


Dr Sally Stowe
Facility Coordinator
Australian National University EM Unit
Research School of Biological Sciences
Box 475
ACT 2601, Canberra, Australia
FAX 06 (0)2 6279 8525






From: Robert Santoianni :      Robert_Santoianni-at-emory.org
Date: Mon, 28 Dec 1998 06:05:05 -0500
Subject: Microtome Service Contracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To Mike D:
In the SE, try QA Support Services 404-627-3715.





From: Michal Jarnik :      M_Jarnik-at-FCCC.edu
Date: Mon, 28 Dec 1998 09:12:15 -0500
Subject: Microtome service

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.
--------------7DCF927DFE885DC8910A00A0
Content-Type: text/plain; charset=us-ascii; x-mac-type="54455854"; x-mac-creator="4D4F5353"
Content-Transfer-Encoding: 7bit

You may also try MOC (914-268-6450), we have been using them for a
couple of years and they are OK.



--
Michal Jarnik, Ph.D.
Fox Chase Cancer Center
Electron Microscopy Facility
7701 Burholme Ave.
Philadelphia PA 19111
Tel. 215-728-5675
Fax 215-728-2412


--------------7DCF927DFE885DC8910A00A0
Content-Type: text/x-vcard; charset=us-ascii;
name="M_Jarnik.vcf"
Content-Transfer-Encoding: 7bit
Content-Description: Card for Michal Jarnik
Content-Disposition: attachment;
filename="M_Jarnik.vcf"

begin:vcard
n:Jarnik;Michal
x-mozilla-html:FALSE
adr:;;;;;;
version:2.1
email;internet:M_Jarnik-at-fccc.edu
x-mozilla-cpt:;3
fn:Jarnik, Michal
end:vcard

--------------7DCF927DFE885DC8910A00A0--






From: Yoo choong geun :      magnetic-at-hymail.hanyang.ac.kr
Date: Tue, 29 Dec 1998 11:08:24 +0900
Subject: Questions for cross-sectional TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


--=====================_914864904==_
Content-Type: text/plain; charset="iso-8859-1"
Content-Transfer-Encoding: quoted-printable


Dear Listserver

Hello.=20
My name is choong keun Yoo and graduate student at hanyang university in
korea.
I have a question for sample preparation in cross-sectional TEM.
i will explain my problems following paragraph.

Specimen size : thickness 20=A7=AD , width 3mm , length 3cm=20
Specimen type : ribbon
Composition : Fe-Co-B-Si

At first, using Si dummy wafers outside of two ribbons along with epoxy, Si
dummy wafers were broken in cutting with diamond saw.
So next step is to use moly and copper tubing kit. (ref. attached file)
but next problem is that specimen's thickness is too slim to fix to moly
and copper kit.=20
And the material putting in the space is so difficult to make and they(
side material and specimen) are different in dimpling and ion milling rate.
So prepered hole is generated in dimpling. (ref. attached file)
So far, i said my problems in experiment

How can i promote my experiment?
What method is good for that case?
Wanna your help...

Thank you for reading.

--=====================_914864904==_
Content-Type: application/octet-stream; name="Crosssection TEM.ppt";
x-mac-type="534C4433"; x-mac-creator="50505433"
Content-Transfer-Encoding: base64
Content-Disposition: attachment; filename="Crosssection TEM.ppt"
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=
--=====================_914864904==_
Content-Type: text/plain; charset="us-ascii"



--=====================_914864904==_--






From: rice-at-mcc.com (Janet Rice)
Date: Tue, 29 Dec 1998 07:58:19 -0500
Subject: Training for SEM work on Semiconductors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can one of you out there on the listserv suggest a training program for
electron microscopy of semiconductors? I'm looking for training in both
sample prep techniques and in SEM technique as well. If we aren't likely to
upgrade equipment at the moment, upgrading skills would not be a bad idea.


Janet Rice
MCC
Senior Member Technical Staff
rice-at-mcc.com
512-338-3266







From: Steve Chapman :      PROTRAIN-at-CompuServe.COM
Date: Wed, 30 Dec 1998 02:59:17 -0500
Subject: EM Training

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

We run courses "in house", in your own laboratory on your own equipment, =
on
all aspects of electron microscopy. Please take a look at our web site f=
or
more information, we are able to tune a course to any requirement. =


Advantages of using your own equipment are that we are able to see
problems that may be due to that particular equipment. AIso imaging on
different systems, different specimen detector geometry in the SEM, may
give considerably different results,

Steve Chapman

Senior Consultant E.M.
Protrain, 16 Hedgerley, Chinnor, Oxford OX9 4TN, England.
Tel & Fax 44 (0)1844 353161
Web Site - http://ourworld.compuserve.com/homepages/protrain
For Consultancy and Courses in Electron Microscopy World Wide





From: LAB. DE MICROSCOPIA ELECTRONICA - FI - UNER :      microsc-at-fi.uner.edu.ar
Date: Wed, 30 Dec 1998 09:00:17 +0000
Subject: microwave processing

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


hi:
I'm looking for tissue processing technics using microwave in SEM and TEM.
any help is welcome

thanks in advance
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D
Fernando D. Balducci
Laboratorio de Microscopia Electr=F3nica
Facultad de Ingenier=EDa - Bioingenieria
Universidad Nacional de Entre Rios
Argentina
e-mail: microsc-at-fi.uner.edu.ar
tel: 54 43 975100
fax: 54 43 975077


=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D





From: kie26-at-ipax.com.au (dyaftxm)
Date: Wed, 30 Dec 1998 20:54:27 +0900
Subject: 'Tis the day after Christmas And something's not right....

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



'Tis the day after Christmas
And something's not right
We’re painfully aware
That our clothes are too tight.

We ate lots of gravy
And pumpkin pie too
Oh, no, not a diet...
But what shall we do?

Now you can lose weight without suffering. Call Tides of Life at
1-888-817-7443 for information on safe, natural, and effective weight
loss products.

Our premier product, Cheat & Eat, contains 1000 mg of chitosan per
tablet along with herbs. This binds to fats and sugars in the
intestinal tract, blocking absorption and passing them out of the body
before you can absorb the calories! Also stops sugar cravings.
Clinical studies show an average weight loss of 4 pounds per week with
no drugs, side effects or starvation diets! We are offering a great
introductory price of $29.50 plus $2.50 shipping for a 6 week supply.
You can buy much weaker chitosan products for $70/bottle but we have
located the best product for you at the best price.

Call 1-888-817-7443 for more information on Cheat & Eat and other
natural weight loss products. Tides of Life has specialized in natural
hormonal products and related health supplements since 1993. Free price
list upon request.

*-*-*-*-*-*-*-*
Removal from lists, please go to http://209.216.69.189/remove/
*-*-*-*-*

*-*
=: 'Tis the day aft





From: Dave_Work-at-student.uml.edu (Dave Work)
Date: Wed, 30 Dec 1998 15:15:01 -0500
Subject: Optical Microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I am looking for information on techniques using optical microscopy to
quantify flaws within hardened concrete and grout samples. I am
currently using Image Pro 3.0 with the Materials Pro module to quantify
images off of a video microscope and am looking for more in-depth
information. Thanks.

Dave






From: Marek Malecki, M.D., Ph.D. :      malecki-at-MACC.WISC.EDU
Date: Wed, 30 Dec 1998 19:51:10 -0600
Subject: Turbomolecular pump based vacuum system for sale.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to sell magnetically levitated turbomolecular pump based - TPU
180, TCM 180, MD4T Balzers - oil-free pumping system ($20420.- original
cost a year ago).
Alternatively, I would be interested in trading this system in for
cryo-pumps. I will be happy to provide more detailed information to or to
discuss offers with those who are interested.
Marek Malecki.




Marek Malecki, M.D., Ph.D.
address: IMR, 1675 Observatory Drive, Madison, WI 53706.
cellular phone: 6084441680.
voice mail: 6082638481.
fax: 6082654076.
email: malecki-at-macc.wisc.edu
http://www.wisc.edu/cesip/
http://www.bocklabs.wisc.edu/imr/integrat/transg.htm







From: Laurie Smith :      lsmith-at-biomail.ucsd.edu
Date: Wed, 30 Dec 1998 18:11:24 -0800
Subject: Job opening for electron microscopist

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


All interested are invited to apply! This announcement can also be found
at:
http://www-biology.ucsd.edu/positions/molGenBioChemCell.html

Manager and Electron Microscopist
Biology Department Electron Microscopy Facility
University of California, San Diego, La Jolla, CA
Job title: Associate Specialist (salary range $42,660-$45,624)

A position is available January 1, 1999, for an individual with extensive
expertise in the preparation and analysis of biological materials for
electron microscopy. A Ph.D. or M.S. degree in the life sciences with
substantial graduate research experience in electron microscopy and a high
level of skill in preparation and analysis of samples by transmission
electron microscopy are required. Experience with the following techniques
is preferred: negative staining and immunogold labelling for TEM,
cryofixation and cryosectioning, scanning electron microscopy, confocal
microscopy, developing and printing of photographic negatives, and
preparation of digital images.

The individual hired will have primary responsibility for all aspects of
the Biology Department Electron Microscopy Facility's operation. Essential
functions include performing electron microscopy on a fee-for-service basis
for a wide range of scientific projects, as well as training and assisting
users of the Facility in sample preparation and analysis. The Manager will
oversee maintenance of equipment and the Facility in general, as well as
record keeping, ordering, and upkeep of supplies.

The facility serves a community of active UCSD and biotech scientists. The
Manager of the EM facility will interact with top EM facilities in the La
Jolla area. Opportunities for grant writing and involvement in growth of
the facility are possible.

Closing date for applications is February 15. Inquiries about the position
may be sent to:lsmith-at-biomail.ucsd.edu. Send applications (CV with names,
addresses, phone and Fax numbers and email addresses of 3 references) to:

Carrie Garnett
Supervisor, Academic Affairs
Department of Biology - 0346
UCSD
9500 Gilman Drive
La Jolla, CA 92093-0346

UCSD is an equal opportunity/affirmative action employer with a strong
institutional commitment to the achievement of diversity among its faculty
and staff.

Laurie G. Smith
Assistant Professor
Biology Dept. 0116
U.C.San Diego
9500 Gilman Drive
La Jolla, CA 92093-0116
email: lsmith-at-biomail.ucsd.edu
telephone: 619-822-2531 (office)
619-822-2558 (lab)
fax: 619-534-7108







From: ricardo :      ricardo-at-ans.com.au
Date: Thu, 31 Dec 1998 17:30:21 +1100
Subject: SEM and beetles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is a multi-part message in MIME format.

------=_NextPart_000_0110_01BE34E3.3F17D6A0
Content-Type: text/plain;
charset="iso-8859-2"
Content-Transfer-Encoding: 7bit

I would like to know if there is any new technique about SEM and beetles,
what is the best way to mount the beetle.

Keep care and be of good cheer.

Regards

Vratislav Richard Eugene Maria John Baptiste
of Bejsak (Bayshark)-Collorado-Mansfeld

Coleoptera - Australia, Tenebrionidae of World
(incl. Lagriinae, Alleculinae)

Temporally home address:
32 Girrawheen Ave.
Kiama NSW 2533
Australia
e-mail: vratislav-at-bigfoot.com
ricardo-at-ans.com.au
(before Ricardo-at-compuserve.com
and ricardo-at-login.cz )

http://www.coleoptera.org
phone : 0414 540 465 (Australia)
+61 414 540 465 (International)

Only after the last tree has been cut down,
only after the last river has been poisoned,
only after the last fish has been caught,
only then will you find that money can not be eaten.'
CREE INDIAN PROPHECY.



------=_NextPart_000_0110_01BE34E3.3F17D6A0
Content-Type: text/x-vcard;
name="Vratislav Richard Eugene Maria John Baptist Bejsak-Collorado-Mansfeld.vcf"
Content-Transfer-Encoding: quoted-printable
Content-Disposition: attachment;
filename="Vratislav Richard Eugene Maria John Baptist Bejsak-Collorado-Mansfeld.vcf"

BEGIN:VCARD
VERSION:2.1
N:Bejsak-Collorado-Mansfeld;Vratislav Richard;Eugene Maria John Baptist
FN:Vratislav Richard Eugene Maria John Baptist Bejsak-Collorado-Mansfeld
ORG:Bayshark Research Laboratory
TITLE:director
NOTE;ENCODING=3DQUOTED-PRINTABLE:Marketing and =
Coaching=3D0D=3D0A=3D0D=3D0ATenebrionidae Orbis and higher taxonomy
TEL;WORK;VOICE:(+61 2) 9319 6380
TEL;CELL;VOICE:(+61 414) 540 465
ADR;WORK:;;29 Edward Street;Darlington, SYDNEY;;NSW 2008;Australie
LABEL;WORK;ENCODING=3DQUOTED-PRINTABLE:29 Edward =
Street=3D0D=3D0ADarlington, SYDNEY NSW 2008=3D0D=3D0AAustralie
ADR;HOME;ENCODING=3DQUOTED-PRINTABLE:;;(temporaly address):=3D0D=3D0A32 =
Girrawheen Ave;KIAMA;;NSW 2533;AUSTRALIA
LABEL;HOME;ENCODING=3DQUOTED-PRINTABLE:(temporaly address):=3D0D=3D0A32 =
Girrawheen Ave=3D0D=3D0AKIAMA NSW 2533=3D0D=3D0AAUSTRAL=3D
IA
URL:
URL:http://www.coleoptera.org
EMAIL;INTERNET:ricardo-at-login.cz
EMAIL;PREF;INTERNET:vratislav-at-bigfoot.com
EMAIL;INTERNET:ricardo-at-ans.com.au
REV:19981231T063007Z
END:VCARD

------=_NextPart_000_0110_01BE34E3.3F17D6A0--






From: Barbara Foster :      mme-at-map.com
Date: Thu, 31 Dec 1998 14:43:34 -0500
Subject: Re: Microtome service contracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mike,


Where are you located?


Barbara Foster

Consortium President

{color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
..Educating microscopists for greater productivity.


{/color} 125 Paridon Street Suite 102 Springfield, MA 01118

PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

Visit our web site { {http://www.MME-Microscopy.com/education}

******************************************************

{bigger} {bigger} MME {/bigger} {/bigger} is America's first national
consortium dedicated to

customized on-site training in all areas of

microscopy, sample preparation, and image analysis {bigger} .

{/bigger}


At 09:59 AM 12/24/98 -0500, MICHAEL DELANNOY wrote:

} ------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

} -----------------------------------------------------------------------.

}

}

} Happy Holidays,

} Does anyone know of a company or small contractor who will

} service Reichert microtomes (other than the more expensive bigger

} companies like Leica). I recall a company called Tek-net in N.J.

} but am unsure. Salutaions,

}

} Mike D

}

}

}

}

}







From: Barbara Foster :      mme-at-map.com
Date: Thu, 31 Dec 1998 14:51:24 -0500
Subject: Re: Training for SEM work on Semiconductors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Janet,


Microscopy/Microscopy Education just added Lisa Montanaro to our staff.
She has nearly a decade of experience in this area, having worked for IBM
and for Dominion. (She was given the "Best of the Best" award at IBM).
Since we specialize in integrating your level of experience, with your
equipment and your application, we can tailor a course specifically to
fit your needs. Our offices will be open again next Monday. Please call
and we can discuss details.


Happy holidays!

Barbara Foster

Consortium President

{color} {param} 0000,8080,0000 {/param} Microscopy/Microscopy Education
..Educating microscopists for greater productivity.


{/color} 125 Paridon Street Suite 102 Springfield, MA 01118

PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

Visit our web site { {http://www.MME-Microscopy.com/education}

******************************************************

{bigger} {bigger} MME {/bigger} {/bigger} is America's first national
consortium dedicated to

customized on-site training in all areas of

microscopy, sample preparation, and image analysis {bigger} .

{/bigger}



At 07:58 AM 12/29/98 -0500, Janet Rice wrote:

} ------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of America


} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

} -----------------------------------------------------------------------.

}

}

} Can one of you out there on the listserv suggest a training program
for

} electron microscopy of semiconductors? I'm looking for training in
both

} sample prep techniques and in SEM technique as well. If we aren't likely
to

} upgrade equipment at the moment, upgrading skills would not be a bad
idea.

}

}

} Janet Rice

} MCC

} Senior Member Technical Staff

} rice-at-mcc.com

} 512-338-3266

}

}

}

}

}







From: Mark Wall :      wall1-at-llnl.gov
Date: Thu, 31 Dec 1998 16:33:53 -0800
Subject: TEM position

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The TEM Service Group in the Chemistry and Materials Science Directorate at
Lawrence Livermore National Laboratory (LLNL)is considering posting a term
position for an Electron Microscopists. The position will be open to all
candidates with a technical or professional degree and 1-2 yrs. of related
work experience.

The primary job function will be to perform a variety of (S)TEM techniques
(as needed) to characterize the microstructure of a wide variety of
materials that are utilized and/or are being developed for internal and
external programs that LLNL supports. Examples include: Multilayered
materials, areo-gels, metallurgical materials, ceramics, etc. Additional
duties will include the use and developement of sample preparation
techniques for TEM. Basic level experience or working knowledge of
conventional, analytical and HREM TEM techniques and associated analysis
techniques is required. Knowledge of materials science, metallurgy,
computers, software, etc. is desired. USA citizenship is preferred.

The position will be a 1yr term, renewable up to 6 years. The salary range
will be =89 $3.5 - 4.5K / month.

Send inquires and/or resumes to:

Mr. Mark A. Wall
Sr. Scientific Assoc.
L-350
7000 East Ave.
C&MS Dir.
Lawrence Livermore National Laboratory
Livermore, CA USA 94551

or

Email WALL1-at-LLNL.GOV

Mr. Mark A. Wall
Sr. Scientific Assoc.
L-350
Chemistry & Materials Science Directorate
Lawrence Livermore National Laboratory
Livermore, CA USA
94550

ph: 925 423-7162
fax: 925 422-6892