From: rfelten-at-Macdermid.com
Date: Wed, 1 Dec 1999 08:45:42 -0500
Subject: SEM/quantitative analysis for P in Ni

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Rick Felten-at-MACDERMID
12/01/99 08:45 AM
I have been measuring the %P in electroless Ni for several years. I have
done some recent work where the numbers that I a getting are 20% lower than
the accepted wet method. I can't imagine what I am missing. I use
standards (Pure Ni and GaP (about 30%P)). I measure the beam current from
the SEM. The current drift is typically around 1%. I even normalize my
data for current drift. My standards have a %rel std dev of about 1%. I
choose several point per sample and standard. Some sample I run un-coated.
In such a run the only thing coated is my manufactured GaP standard, that
has 200A of carbon. It hasn't been re-polished in 6 years, but that should
not give higher P cts in my std. I have tried manually performing my
background subtraction subtracting around the P peak, with only a 3%
improvement. I get an analytical total typically 95-100% (%P + %Ni). This
should be a sanity check for precision and accuracy. I collect data on my
standards each time I run a new sample so tilt or detector window
condensation should have the same effect. The range for %P recently has
been 5-8% which gives a healthy peak. What could I be missing?
Any help would be appreciated.
Thanks
Ric







From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Wed, 01 Dec 1999 10:02:05 -0600
Subject: Re: SEM/quantitative analysis for P in Ni

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I don't know much about your sample or standard, but I would wonder if
there is any heterogeneity in the distribution of P in either your sample
or standard? I would hope the GaP is quite stable, but is there any chance
of formation of a P-rich layer at or near the surface? You aren't probing
very deep. Also, how does the P occur in the nickel. Is it present in sold
solution only or in inclusions of some kind?

You say your results are 20% low for P. Is that relatively speaking? If so,
then it may be that you having standard problems. The 2% or so that you are
missing would help your analytical totals if you can find it.

Some quick thoughts.
Warren S.

At 08:45 AM 12/1/1999 -0500, you wrote:
} Rick Felten-at-MACDERMID
} 12/01/99 08:45 AM
} I have been measuring the %P in electroless Ni for several years. I have
} done some recent work where the numbers that I a getting are 20% lower than
} the accepted wet method. I can't imagine what I am missing. I use
} standards (Pure Ni and GaP (about 30%P)). I measure the beam current from
} the SEM. The current drift is typically around 1%. I even normalize my
} data for current drift. My standards have a %rel std dev of about 1%. I
} choose several point per sample and standard. Some sample I run un-coated.
} In such a run the only thing coated is my manufactured GaP standard, that
} has 200A of carbon. It hasn't been re-polished in 6 years, but that should
} not give higher P cts in my std. I have tried manually performing my
} background subtraction subtracting around the P peak, with only a 3%
} improvement. I get an analytical total typically 95-100% (%P + %Ni). This
} should be a sanity check for precision and accuracy. I collect data on my
} standards each time I run a new sample so tilt or detector window
} condensation should have the same effect. The range for %P recently has
} been 5-8% which gives a healthy peak. What could I be missing?
} Any help would be appreciated.
} Thanks
} Ric






From: Hans Brinkies :      hbrinkies-at-lucy.cc.swin.edu.au
Date: Thu, 2 Dec 1999 08:40:37 +0000
Subject: X-Ray Spectrometer

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Again some items for free, that is, if you can arrange pick up.

I am disposing of a complete ETEC Autospec WD X-ray Spectrometer.
The spectrometer with crystals and separate electronic console
containing:
Detector Control Crystal Select (LiF, PET, LOD, RAP),
Spec.Pos.Control,
Scaler Timer,
Amplifier -PHA'
Linear Ratemeter.
This WDS was originally attached to our ETEC Autoscan SEM, which
(just a point of general interest) is still working satisfactory
after 26 years of usage.

Cheers

Hans Brinkies

Hans G Brinkies
Senior Lecturer
Swinburne, University of Technology
School of Engineering and Science
Electron Microscopy
Hawthorn, 3122, Melbourne - Australia





From: Marti, Jordi :      Jordi.Marti-at-alliedsignal.com
Date: Thu, 2 Dec 1999 11:09:59 -0700
Subject: Glass Interface

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello:

I have a glass (leaded glass) tube, about 0.5 cm in diameter. The inside
wall of the tube has been leached leaving behind a porous silica network
with pores about 500 nm in size. The thickness of the leached layer is
estimated to be about 400 nm. We would like to prepare cross sections to
look at the glass/leached layer interface and to be able to actually measure
the thickness of the leached layer. I'm wondering if anyone has suggestions
or experience on preparing samples such as these for TEM ?. Tripod
polishing is a possibility , but the leached layer might not survive the
polishing. Has anyone tried microtomy with this type samples?.

Suggestions are welcomed.

Thanks

Jordi Marti






From: Soumitra Ghoshroy :      ghoshroy-at-nmsu.edu
Date: Thu, 02 Dec 1999 11:40:51 -0700
Subject: Microscopy book

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear All,

Can anyone suggest a good text book on light microscopy, principles,
techniques etc ? It will be great if the particular book includes chapters
on fluorescence microscopy. Otherwise you can also give me the title for a
separate book on fluorescence microscopy.

Thanks in advance.

Soumitra



Soumitra Ghoshroy Ph.D.
Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003-8001
Tel: 505-646-1531/3600
Fax: 505-646-5665
e-mail: ghoshroy-at-nmsu.edu





From: Marti, Jordi :      Jordi.Marti-at-alliedsignal.com
Date: Thu, 2 Dec 1999 14:12:24 -0700
Subject: Reichert Cryo-Microtome Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello:

We have a Reichert Ultracut E with an FC4D cryo unit. Lately I have been
under the impression that the reservoir refilling cycle is taking much
longer than it used to.

I was wondering if others with the same unit could give me their
impression as to whether or not the behavior I describe bellow seems normal.

When I first start the unit with a filled LN dewar, the first two reservoir
refilling cycles(by this I mean the time it takes to go from only one
indicating green light on to all green lights on) take only a few minutes (
maybe five minutes), but after that the time to refill the reservoir
decreases considerably. After an hour or so of using the unit, with the
temperature (knife and sample) stabilized at about -100C, it takes close to
25 min. to refill the reservoir. I also notice that the temperature of the
knife and sample changes from a low of -100C (with the reservoir full) to
-94 with the reservoir empty (i.e. when the refill just starts). If someone
would let me know if this seems normal I would appreciate it.

Thanks

Jordi Marti





From: Jos=?ISO-8859-1?B?6SBVbGzh?=n Serrano :      jullan-at-mixcoac.upmx.mx
Date: Thu, 02 Dec 1999 16:57:49 -0600
Subject: Epifluorescence

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everyone!

Please respond directly to the e-dress below.

I am looking for an epifluorescence equipment to fit our OM Zeiss
Universal.
Would anyone care to trade or sell those parts I need?
- the 50 W lamp illuminator equipment (or 100 W)
- the epifluorescence' nosepiece (whitout the Neofluar objectives).
- not indispensable: exciter and barrier filters (BG12 & 53/44; BG3 & 50/44)

We would appreciate any info on the track record of these OM products.
My request to the Carl Zeiss web site was processed, but without result
after tree weeks.
Thanks in advance

JOSE ULLAN E-mail: jullan-at-mixcoac.upmx.mx
-----------------------------------
Dept. ANATOMIA
E. de MEDICINA
Univ. Panamericana http://www.mixcoac.upmx.mx





From: Keith Ryan :      KPR-at-wpo.nerc.ac.uk
Date: Fri, 03 Dec 1999 10:41:50 +0000
Subject: Re: Reichert Cryo-Microtome Question

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


If I recall the legends correctly it was some sort of 'growth'
Ed Sharpe


The Microscopy ListServer -- Sponsor: The Microscopy Society of America


}
} I am wondering if this is the infamous 'CD Rot' condition that was
} prophesied at the dawn of the CD era.
}
} Ed Sharpe archivist for SMECC
}

Ed,

OK I'll bite. Tell us more about this 'CD Rot'. I think not leaving them
lying around exposed to direct sunlight or even strong daylight for years
on end would be a reasonable precaution to take.

(Does anybody know of a suitable fixitive ;)





Arthur Day, Electron Microscope Unit Phone: 61-2-9717-3457
Ansto Materials Division Fax: 61-2-9543-7179
PMB 1, Menai (Sydney), NSW, 2234 Email: ard-at-ansto.gov.au
Australia www: http://www.ansto.gov.au/





----------------------- Headers --------------------------------
Return-Path: {Microscopy-request-at-sparc5.microscopy.com}
Received: from rly-zd03.mx.aol.com (rly-zd03.mail.aol.com [172.31.33.227])
by air-zd01.mail.aol.com (vx) with ESMTP; Thu, 02 Dec 1999 18:52:40 -0500
Received: from Sparc5.Microscopy.Com ([206.69.208.10]) by
rly-zd03.mx.aol.com (v65.4) with ESMTP; Thu, 02 Dec 1999 18:52:28 -0500
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11) id
TAA28170 for dist-Microscopy; Tue, 30 Nov 1999 19:12:58 -0600
Received: from no_more_spam.com (Sparc5 [206.69.208.10]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id TAA28167 for
"MicroscopyFilteredEmail-at-msa.microscopy.com"; Tue, 30 Nov 1999 19:12:26 -0600
Received: from tachyon.gw.ansto.gov.au (anstogw.gw.ansto.gov.au
[137.157.8.253]) by Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id
TAA28160 for {Microscopy-at-sparc5.microscopy.com} ; Tue, 30 Nov 1999 19:12:04
-0600
Received: by tachyon.gw.ansto.gov.au; id MAA23463; Wed, 1 Dec 1999 12:11:19
+1100 (EST)
Received: from hadron.ansto.gov.au(137.157.13.219) by tachyon.gw.ansto.gov.au
via smap (4.1)
id xma023305; Wed, 1 Dec 99 12:10:39 +1100
Received: from [137.157.95.82] (arthur.amat.ansto.gov.au [137.157.95.82])
by hadron.ansto.gov.au (8.9.3/8.9.3) with SMTP id MAA09324
for {Microscopy-at-sparc5.microscopy.com} ; Wed, 1 Dec 1999 12:10:39 +1100
(EST)
X-Sender: ard-at-hadron.ansto.gov.au
Message-Id: {v02140b03b46a2352ebc3-at-[137.157.95.82]}
Mime-Version: 1.0
Content-Type: text/plain; charset="us-ascii"


} From: Chris Jeffree {cjeffree-at-srv0.bio.ed.ac.uk}
To: "Marti, Jordi" {Jordi.Marti-at-alliedsignal.com}


I have some experience with the original FC4. How about checking the
state of the seal between the pump and the dewar? Ours cracked and
was replaced with an O-ring - not ideal but it works - maybe you are
losing filling pressure? The electronics should either work or fail -
that is just my guess.

Also, maybe check the hose pipe connections for wear or damage -
again with pressure retention in mind.

Keith Ryan
_________

_______________________________
Dr. Keith Ryan
Marine Biological Association of the UK
The Laboratory
Citadel Hill
Plymouth
Devon PL1 2PB
England

Tel. ++44 (0)1752 633249
Tel. ++44 (0)1752 633279
The 279 number has an answering machine

Fax ++44 (0)1752 633102

e-mail: kpr-at-wpo.nerc.ac.uk





From: Ken Converse :      qualityimages-at-netrax.net
Date: Fri, 03 Dec 1999 07:48:10 -0800
Subject: Re: Glass Interface

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Marti, Jordi wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Hello:
}
} I have a glass (leaded glass) tube, about 0.5 cm in diameter. The inside
} wall of the tube has been leached leaving behind a porous silica network
} with pores about 500 nm in size. The thickness of the leached layer is
} estimated to be about 400 nm. We would like to prepare cross sections to
} look at the glass/leached layer interface and to be able to actually measure
} the thickness of the leached layer. I'm wondering if anyone has suggestions
} or experience on preparing samples such as these for TEM ?. Tripod
} polishing is a possibility , but the leached layer might not survive the
} polishing. Has anyone tried microtomy with this type samples?.
}
} Suggestions are welcomed.
}
} Thanks
}
} Jordi Marti


Jordi,
With a pore size of 500nm, have you considered looking at a fracture
mounted such that the original tube would be vertical and using an SEM?
I would try coating with carbon, followed by gold (or Au/Pd), and use as
low a kV as possible.

Ken Converse
Quality Images
third party SEM service
Delta, PA





From: Ken Converse :      qualityimages-at-netrax.net
Date: Fri, 03 Dec 1999 07:58:34 -0800
Subject: Re: ETEC Autospec

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hans Brinkies wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Again some items for free, that is, if you can arrange pick up.
}
} I am disposing of a complete ETEC Autospec WD X-ray Spectrometer, ie.
} the spectrometer with crystals and separate electronic
} operational console containing; Detector Control
} Crystal Select (LiF, PET, LOD, RAP)
} Spec.Pos.Control
} Scaler Timer
} Amplifier -PHA
} Linear Ratemeter.
} This WDS was originally attached to our ETEC Autoscan SEM, which
} (just a point of general interest) is still working satisfactory
} after 26 years of usage.
}
} Cheers
}
} Hans Brinkies
}
} Hans G Brinkies
} Senior Lecturer
} Swinburne, University of Technology
} School of Engineering and Science
} Electron Microscopy
} Hawthorn, 3122, Melbourne - Australia



Hans,
I'm not surprised at all that your Autoscan is still running fine. I'm
still servicing #17 (vintage '71 or early '72) that has been abused by
hundreds of students. And you can't beat the camera system unless you
go to a 16mB file size. I have a number of Autoscans on their second or
third owner. Most of the original operators only got rid of them
because their management or bean-counters told them they had to. The
"new" ones are only about 17 years old.

Keep a good system running!

Ken Converse
Quality Images
third party SEM service
Delta, PA





From: MICHAEL DELANNOY :      delannoy-at-mail.jhmi.edu
Date: Fri, 3 Dec 1999 09:44:03 -0500 (EST)
Subject: IEM ribosomal associated proteins

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,
Does anyone have experience with immuno em of ribosomal associated proteins,
fixation to embedding, pre or post embed label. The If works but we are not
sure that our standard IEM protocol (post embed label/ LR White) will work.
We would appreciate advise before we dive in . Thanks

Mike D







From: James Martin :      James.S.Martin-at-williams.edu
Date: Fri, 03 Dec 1999 09:45:25 -0500 (EST)
Subject: additional info about spotty CD-Rs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


First, thanks to all who replied to my inquiry about "spotty CD-Rs."
One respondent described a similar problem with CD-Rs (CDQ-74BN) from the
same manufacturer. My disks are all CDQ-74CN.

Since posting the inquiry I've inspected CD-Rs written in 9/1999, and one
of these showed crystalline spots - which developed since the disk was
written and kept in dark storage at moderate T and RH.

This morning I sampled and analyzed the crystalline material using FTIR
microscopy. The crystalline sample spectrum compares with reference
spectra for bisphenyl-A. The CD-R surface spectrum compares with
reference spectra for polycarbonate, bisphenyl-A. I am not knowledgable
about the manufacture and composition of CD-Rs, but it seems that
bisphenyl-A, or a similar compound, is recrystallizing on the CD-R
surface.

Can anyone provide more enlightened interpretation of this data?

Jamie

-------------------------------------------------------------------------

James Martin
Director of Analytical Services & Research
Williamstown Art Conservation Center
james_martin.tripod.com/dasr.htm

Research Scientist in Chemistry
Williams College
james_martin.tripod.com/williams.htm

*** Please don't send e-mail attachments. Cut-and-paste text into the
body of an e-mail message. ***






From: Brian McIntyre :      mcintyre-at-optics.rochester.edu
Date: Fri, 3 Dec 1999 10:29:49 -0500
Subject: Re: Glass Interface

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


hi jordi-

i've microtomed some glass/multilayer coating surfaces. this tends to make
a mess of the knife (try a used diamond). you need to pick up the "pieces"
on a filmed grid and hunt for a good area. the pieces are really fractured
instead of cut so there is a lot of variation in them. its really not too
difficult, but it may end up being costly.

good luck!
b-
************************************************************
}
}
} Hello:
}
} I have a glass (leaded glass) tube, about 0.5 cm in diameter. The inside
} wall of the tube has been leached leaving behind a porous silica network
} with pores about 500 nm in size. The thickness of the leached layer is
} estimated to be about 400 nm. We would like to prepare cross sections to
} look at the glass/leached layer interface and to be able to actually measure
} the thickness of the leached layer. I'm wondering if anyone has suggestions
} or experience on preparing samples such as these for TEM ?. Tripod
} polishing is a possibility , but the leached layer might not survive the
} polishing. Has anyone tried microtomy with this type samples?.
}
} Suggestions are welcomed.
}
} Thanks
}
} Jordi Marti








From: Sobocinski, Gregg :      Gregg.Sobocinski-at-WL.com
Date: Fri, 3 Dec 1999 10:35:14 -0500
Subject: additional info about spotty CD-Rs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Pardon me if someone has already addressed this inquiry, but what did the
manufacturer say about the "spotty CD-Rs"??? They should at least have a
theory about what you are seeing, and might save us some time and
speculation, here. I would recommend you speak to them before running any
major investigations of your own.

Gregg Sobocinski
Parke-Davis Pharmaceuticals
Ann Arbor, Michigan
USA
Gregg.Sobocinski-at-wl.com


-----Original Message-----
} From: James Martin [mailto:James.S.Martin-at-williams.edu]
Sent: Friday, December 03, 1999 9:45 AM
To: MSA listserver


First, thanks to all who replied to my inquiry about "spotty CD-Rs."
One respondent described a similar problem with CD-Rs (CDQ-74BN) from the
same manufacturer. My disks are all CDQ-74CN.

Since posting the inquiry I've inspected CD-Rs written in 9/1999, and one
of these showed crystalline spots - which developed since the disk was
written and kept in dark storage at moderate T and RH.

This morning I sampled and analyzed the crystalline material using FTIR
microscopy. The crystalline sample spectrum compares with reference
spectra for bisphenyl-A. The CD-R surface spectrum compares with
reference spectra for polycarbonate, bisphenyl-A. I am not knowledgable
about the manufacture and composition of CD-Rs, but it seems that
bisphenyl-A, or a similar compound, is recrystallizing on the CD-R
surface.

Can anyone provide more enlightened interpretation of this data?

Jamie

-------------------------------------------------------------------------

James Martin
Director of Analytical Services & Research
Williamstown Art Conservation Center
james_martin.tripod.com/dasr.htm

Research Scientist in Chemistry
Williams College
james_martin.tripod.com/williams.htm

*** Please don't send e-mail attachments. Cut-and-paste text into the
body of an e-mail message. ***






From: James Martin :      James.S.Martin-at-williams.edu
Date: Fri, 03 Dec 1999 10:57:36 -0500 (EST)
Subject: RE: additional info about spotty CD-Rs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Fri, 3 Dec 1999, Sobocinski, Gregg wrote:

} Pardon me if someone has already addressed this inquiry, but what did
} the manufacturer say about the "spotty CD-Rs"??? They should at least
} have a theory about what you are seeing, and might save us some time and
} speculation, here. I would recommend you speak to them before running
} any major investigations of your own.

Repeated calls to the manufacturer ended in a voice mail tree. I have
reported the problem, but have not received a reply.

What has been an inconvenience for me, may be a more serious problem for
other list members who use CD-R but are unaware of this potential
deterioration. Empirical observations showed that crystals can form after
CD-Rs are written and stored, rendering data inaccessible. My series of
analyses took little time - the beauty of FTIR microscopy - but suggested
an experimental composition and source of the crystals.

So ... just be aware.

Jamie

On Fri, 3 Dec 1999, Sobocinski, Gregg wrote:

} Pardon me if someone has already addressed this inquiry, but what did the
} manufacturer say about the "spotty CD-Rs"??? They should at least have a
} theory about what you are seeing, and might save us some time and
} speculation, here. I would recommend you speak to them before running any
} major investigations of your own.
}
} Gregg Sobocinski
} Parke-Davis Pharmaceuticals
} Ann Arbor, Michigan
} USA
} Gregg.Sobocinski-at-wl.com
}
}
} -----Original Message-----
} From: James Martin [mailto:James.S.Martin-at-williams.edu]
} Sent: Friday, December 03, 1999 9:45 AM
} To: MSA listserver
} Subject: additional info about spotty CD-Rs
}
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} First, thanks to all who replied to my inquiry about "spotty CD-Rs."
} One respondent described a similar problem with CD-Rs (CDQ-74BN) from the
} same manufacturer. My disks are all CDQ-74CN.
}
} Since posting the inquiry I've inspected CD-Rs written in 9/1999, and one
} of these showed crystalline spots - which developed since the disk was
} written and kept in dark storage at moderate T and RH.
}
} This morning I sampled and analyzed the crystalline material using FTIR
} microscopy. The crystalline sample spectrum compares with reference
} spectra for bisphenyl-A. The CD-R surface spectrum compares with
} reference spectra for polycarbonate, bisphenyl-A. I am not knowledgable
} about the manufacture and composition of CD-Rs, but it seems that
} bisphenyl-A, or a similar compound, is recrystallizing on the CD-R
} surface.
}
} Can anyone provide more enlightened interpretation of this data?
}
} Jamie
}
} -------------------------------------------------------------------------
}
} James Martin
} Director of Analytical Services & Research
} Williamstown Art Conservation Center
} james_martin.tripod.com/dasr.htm
}
} Research Scientist in Chemistry
} Williams College
} james_martin.tripod.com/williams.htm
}
} *** Please don't send e-mail attachments. Cut-and-paste text into the
} body of an e-mail message. ***
}
}






From: Rujida Leepipattanawit :      leepipat-at-pilot.msu.edu
Date: Fri, 3 Dec 1999 11:15:18 -0500 (EST)
Subject: need any suggestion

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,

I am working on the PET/PEN blends.
Now I try to get the microsturcture images of this Polymer blends by TEM.
However, From the frist experiment, I did not get very good images.
Please suggest about "the sample Preparation".

Again I am an unexperienced and very new in this area.


Thank you
Rujida






From: tschwach :      tschwach-at-barrishind.com
Date: Fri, 3 Dec 1999 10:25:31 -0600
Subject: Re: matrigel penetration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I've used matrigel as well as collagen to coat culture dish and coverslip
surfaces so my tissue culture would grow and proceeded to have no problems
fixing them for EM. It would help to have more information about the
problems you are having.
Tina






From: Barbara Foster :      mme-at-map.com
Date: Fri, 03 Dec 1999 11:53:08 -0500
Subject: Re: Microscopy book

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Soumitra,

"Optimizing Light Microscopy for Biological and Clinical Laboratories" was
written just for this purpose. Jim Pawley has been using it at both U. WI
and UBC; Dave Knecht uses it at UConn. Has learning objectives at the
beginning of each chapter as well as short quizzes at the end. Covers both
basic principles and has a detailed section on fluorescence... even a bit
on confocal.

Classroom discounts are also available. Please see
MME-Microscopy.com/education for further details.

CAVEAT: MME does have a commercial interest in this book.

Best regards,
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************

At 11:40 AM 12/2/99 -0700, Soumitra Ghoshroy wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America






From: G. Macdonald :      glenmac-at-u.washington.edu
Date: Fri, 3 Dec 1999 11:11:56 -0800 (PST)
Subject: RE: additional info about spotty CD-Rs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Could you please let us know the manufacturer? We've got 3 or 4 different
brands over the few years we've been burning CDs. How old were these
disks?

Regards,
Glen


Glen MacDonald
Research Scientist
Hearing Research Laboratories of the
Virginia Merrill Bloedel Hearing Research Center
Box 35-7923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156
glenmac-at-u.washington.edu









From: Malis, Tom :      malis-at-nrcan.gc.ca
Date: Fri, 3 Dec 1999 14:20:21 -0500
Subject: RE: Glass Interface

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You're showing the life sciences general view of sectioning such material,
Chris. In what is loosely termed 'materials science' microtomy, this would
not be that great a challenge if all that is desired is the thickness. I
don't want to steal Phil Swab's thunder, but in case he doesn't see this,
the trick would be to break the tube and embed (with good infiltration re
the porosity!) a pointed shard with sharp end towards the block face.
(Treating the glass with a silane coupling agent - Dow Corning Z-6040 to
enhance adhesion would help). Then section away. Phil has used this method
for all manner of glasses, plus thin film coatings such as - hold your
breath - boron nitride on diamond on silicon substrate (see Microscopy
Research and Technique, vol. 31, p. 308, 1995).

I agree with your FIB comments whole-heartedly, but would also not be
surprised if tripod polishing worked. Again, good infiltration would be the
key, I suspect, though I am no expert on tripoding.

Tom Malis

----------
From: Chris Jeffree [SMTP:cjeffree-at-srv0.bio.ed.ac.uk]
Sent: December 03, 1999 4:43 AM
To: microscopy-at-Sparc5.Microscopy.Com
Subject: Re: Glass Interface


------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html

-----------------------------------------------------------------------.


} From: Chris Jeffree {cjeffree-at-srv0.bio.ed.ac.uk}
To: "Marti, Jordi" {Jordi.Marti-at-alliedsignal.com}
Subject: Re: Glass Interface
Copies to: microsopy-at-sparc5.microscopy.com
Send reply to: c.jeffree-at-ed.ac.uk
Date sent: Fri, 3 Dec 1999 09:37:59 +0000

Dear Jordi
Conventional microtomy won't work on this kind of hard brittle
friable material.
The most effective method to section hard, brittle material such as
glass, silica, ceramics, for TEM is to use a FIB. FIB (focussed ion
beam) sections are made routinely by the microelectronics
industry to examine the structure of their silicon devices and the
interfaces between the fabricated layers. I don't know who offers a
service of this kind in the US but I could put you in touch with a
colleague in UK with whom we collaborate in providing a service for
sectioning of devices. Alternatively, I could arrange the whole job

for you, if that would be useful, and I would be willing to quote
you
for the work.

An alternative to TEM would be to fracture the tube, or cut and
polish it to produce a specimen suitable for SEM. Field emission
SEM at low kV is especially good for this kind of porous dielectric
material, which will be difficult to coat with a continuous
conductive
layer.
Yours sincerely
Chris Jeffree
}
} Hello:
}
} I have a glass (leaded glass) tube, about 0.5 cm in diameter.
The inside
} wall of the tube has been leached leaving behind a porous silica
network
} with pores about 500 nm in size. The thickness of the leached
layer is
} estimated to be about 400 nm. We would like to prepare cross
sections to
} look at the glass/leached layer interface and to be able to
actually measure
} the thickness of the leached layer. I'm wondering if anyone has
suggestions
} or experience on preparing samples such as these for TEM ?.
Tripod
} polishing is a possibility , but the leached layer might not
survive the
} polishing. Has anyone tried microtomy with this type samples?.
}
} Suggestions are welcomed.
}
} Thanks
}
} Jordi Marti


=====================================================================
DR CHRIS JEFFREE
BIOSEM - BIOLOGICAL SCIENCES EM FACILITY
UNIVERSITY OF EDINBURGH
Daniel Rutherford Building
King's Buildings, Mayfield Road
EDINBURGH, EH9 3JH, Scotland, UK
Tel. #44 131 650 5345
FAX. #44 131 650 6563
Mobile 0410 585 401
email c.jeffree-at-ed.ac.uk
SEM / TEM bookings sem-at-ed.ac.uk

=====================================================================






From: Scott Robinson :      sjrobin-at-itg.uiuc.edu
Date: Fri, 03 Dec 1999 13:08:36 -0600
Subject: automated TEM negative processor?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi You All--

I've spoken with one or two persons about this but have not collected
specifics. Can someone out there recommend (or discourage me from
attempting to buy) an automated TEM negative processor? We'd like to be
able to somehow take a stack of exposed 3.25 x 4-in. negatives (Kodak
SO-163), pop them into a machine, and have them come out developed, dry,
and 'archival.'

Please let me know what's out there.

Thank you

Scott Robinson (sjrobin-at-uiuc.edu)
Imaging Technology Group
Beckman Institute
University of Illinois at Urbana-Champaign







From: rnessler :      rnessler-at-emiris.iaf.uiowa.edu
Date: Fri, 3 Dec 1999 14:17:41 -0600 (CST)
Subject: Re: matrigel penetration

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


On Fri, 3 Dec 1999, tschwach wrote:}
}
} I've used matrigel as well as collagen to coat culture dish and coverslip
} surfaces so my tissue culture would grow and proceeded to have no problems
} fixing them for EM. It would help to have more information about the
} problems you are having.
} Tina
}
}
Tina,
I spoke with Kathy about the matrigel, so I think I can relate the
concerns. I have had samples where the cells were seeded in a layer of
matrigel. The thickness of the layer was somewhere between 3-5mm, and it
was in a plastic multi-well plate. Between the thickness of the matrigel, and
the fact that plastic friendly dehydrating solvents had to be used,
infiltration was less than ideal. Removing the matrigel from the plate
was not an option. Now, after the clarification, has anyone had good
success getting such a sample embeded?
Randy


Randy Nessler rnessler-at-emiris.iaf.uiowa.edu
Views expressed are my own.






From: Sally Shrom :      sally-at-retina.anatomy.upenn.edu
Date: Fri, 3 Dec 1999 16:35:41 -0500 (EST)
Subject: job opening TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Job opening:
Electron Microscopy Technician. TEM experience required: ultra-sectioning,
operation of TEM, darkroom.

Send resumes to:

Dr. Peter Sterling
123 Anatomy/Chemistry BLDG
University of Pennsylvania
Philadelphia, PA 19104-6058

FAX # 215-573-8093

E-mail: peter-at-retina.anatomy.upenn.edu






From: Rosemary Walsh :      rw9-at-psu.edu
Date: Fri, 3 Dec 1999 19:26:38 -0400
Subject: Protocol for methacrylate removal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listers,
Does anyone have a protocol to remove methacrylate
plastic from tissue sections using 1-acetoxy-2-methoxyethane?
Rosemary

Rosemary Walsh, Manager
The Electron Microscope Facility for the Life Sciences,
A Shared Technology Facility, The Life Sciences Consortium
1 South FrearLab
Penn State University
University Park, PA 16802
(814) 865-0212
rw9-at-psu.edu
http://www.lsc.psu.edu/stf/em/home.html







From: James Pawley :      jbpawley-at-facstaff.wisc.edu
Date: Sat, 4 Dec 1999 11:39:08 +1100
Subject: 5th Live-cell Course at UBC: June 19-29

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all,

I am pleased to announce that the Live-cell Course continues to grow and
prosper. We have worked with over 100 students from 22 contries over the
past 4 years and the need seems to keep on growing.

3D Microscopy of Living Cells is IN!

We hope that those with a strong interest in this topic will be able to
join us in UBC next summer.

Basic info is below but you can get the entire brochure at

http://www.cs.ubc.ca/spider/ladic/course/brochure.htm

Cheers,

Jim P.

=46ifth Annual INTERNATIONAL 10-Day Short Course on

3D Microscopy of Living Cells
June 19 - 29, 2000



=46ourth, Post-course Workshop on

3D Image Processing,
July 1 - 3, 2000



Organized by Prof. James Pawley,
(University of Wisconsin-Madison)
(SEE SABBATICAL ADDRESS AT END OF MESSAGE)


in association with Dr. Elaine Humphrey
UBC BioSciences Microscopy Facility:
University of British Columbia
Vancouver, BC, Canada



DATES

Applications must be received by March 1, 2000
Deposit due April 15, 2000
Registration 5:00 - 7:00 PM Sunday, June 18, 2000
=46irst Lecture 7:30 PM Sunday, June 18, 2000
Live-cell Course ends, noon Thursday, June 29, 2000


APPLICATIONS DUE BY MARCH 1, 2000


More information at :
http://www.cs.ubc.ca/spider/ladic/course/bulletin.html

or

REGARDING COURSE ORGANIZATION:


Prof. James B. Pawley, (on Sabbatical)
Room LG 10, Madsen Building, F-09,
University of Sydney, NSW, 2006
Australia

Ph. 61-2-9351-7548/2351
=46AX 61-2-9351-7682
JBPAWLEY-at-FACSTAFF.WISC.EDU

REGARDING APPLICATIONS

Dr. Elaine Humphrey,
Biosciences EM Facility
Biosciences Building
Univ. of British Columbia
6270 University Blvd.
Vancouver, BC, V6T-1Z4




THE PURPOSE OF THE COURSE

Modern methods of 3D light microscopy promise a revolutionary improvement
in our ability to view living cells. To help convert this promise to
reality for a wider selection of biological scientists, an intensive
eleven-day residential course concentrating on all aspects of 3D Microscopy
of Living Cells will be held at the University of British Columbia, in June
of 2000. The course includes 4 days on 2D techniques, 6 days of 3D
techniques and a summary presentation. It covers basic microscopy to the
highest level confocal microscopy. (A half-day Pre-course is offered for
any who may need to brush up on basic optics!).

Topics include:
o Quantitative 2D light microscopy
o How to keep your cells alive
o 3D imaging in confocal microscopy
o Widefield/deconvolution techniques
o Two-photon excitation microscopy
o Fluorescent and backscattered light signals
o Dye design, characteristics and use
o Pixelation: The Nyquist Criterion
o Lasers and laser tweezers
o Objectives and aberrations
o Scanning-systems: AODs and mirrors
o Optimal pinhole size/photon efficiency
o Detectors: operation and performance
o Video-rate confocal imaging
o Measuring ion concentrations
o Display and measurement of 3D data

Morning lecture/demonstrations lead to hands-on laboratory exercises each
afternoon that will utilize most of the commercial instruments currently
available for 3D microscopic imaging. Students will work in groups of 3 or
4 throughout the discussion and laboratory sessions, and may complete a
live-cell 3D study on their own specimens. In the first four years, over
100 students from 22 countries have attended. Last year, 11 separate, 3D
microscopical workstations were available for student use under the
supervision of a 17-member international faculty. We expect to have even
more workstations in 2000. Including manufacturers representatives, the
teacher/ student ratio will be almost 2:1.

INTERNATIONAL FACULTY

o Stephen Adams University of California-SD
o Ping Chin Cheng SUNY, Buffalo
o Rachel Errington University of Oxford
o Stephan Hell Max Planck Institute, Goettingen
o Ted Inou=E9 Universal Imaging, PA
o Iain Johnson Molecular Probes, OR
o Larry Keenan Cell Robotics, NM
o Ernst Keller Carl Zeiss, Thornwood, NY
o Felix Margadant University of Sydney
o Tim Murphy University of British Columbia
o Sigrid Myrdal Multidimensional Imaging, WA
o Jim Pawley University of Wisconsin-Madison
o Steve Potter California Institute of Technology
o Jason Swedlow University of Aberdeen
o Michael Weis Agriculture Canada
o Nick White Oxford University

TUITION

Course tuition is $2,150 US and includes lunches. On receipt of 50%
deposit, students will receive preliminary group assignments and the
textbook, Handbook of Biological Confocal Microscopy, (Plenum, 1995). The
tuition fee includes the textbook, a binder of handouts, and tickets for
the Opening Reception, the Manufacturer's Reception and the Beach Party.
Accommodations and meals other than lunch are not included in the tuition
fee. The Pre-course is $100 US.

APPLICATIONS

Applicants will complete a questionnaire to assess knowledge level and
field of interest. Enrollment is limited to about 24 participants.
Selection will be made on the basis of background and perceived need.
Those without previous LM experience will be provided with basic texts on
request to read before the course begins and are encouraged to take the
Pre-course on the afternoon of June 18.


Application forms, and other course information from this and past years,
can be downloaded from the WWW site at:

http://www.cs.ubc.ca/spider/ladic/course/bulletin.html

or obtained from:

Dr. Elaine Humphrey,
Biosciences EM Facility
Biosciences Building
Univ. of British Columbia
6270 University Blvd.
Vancouver, BC, V6T-1Z4
Phone: 1-604.822-3354
FAX: 1-604.265-5315
Email: ech-at-unixg.ubc.ca.


Application deadlines:

Application forms are due by March 1, 2000!
Deadlines are early to facilitate setting up groups. Successful applicants
will be notified by April 1, and a deposit of 50% must be received by April
15, 2000 to reserve your position. In general, deposit refunds are only be
possible if your position can be filled from the waiting list. The
remainder of the fees is due at registration.

DATES

Applications must be received by March 1, 2000
Deposit due April 15 2000
Registration 5:00 - 7:00 PM Sunday, June 18, 2000
=46irst Lecture 7:30 PM Sunday, June 18, 2000
Live-cell Course ends, noon Thursday, June 29, 2000


TEACHING PHILOSOPHY

As befits teaching in an area at the boundary of "what is now known,"
lecturers have been chosen based on their expertise as scientists working
in the field rather than because they all agree. They are encouraged more
to be provocative than to be prosaic. Students should expect discussion in
areas where differences of opinion exist.

Prior to the course, students will be organized into groups and encouraged
to communicate by email/phone, about the "Living-cell" group projects that
they will pursue during the course and that will be presented to the class
on the last day of the course. It has been found that group interactions
make best use of students' prior experience and can be very effective in
teaching the practical skills covered in a hands-on course of this type.

ARRANGEMENTS FOR LIVE SAMPLES

Students must contact Dr. Elaine Humphrey to make necessary arrangements
for the transport and maintenance of cell lines etc. needed for their
projects. Organisms linked in any way with human disease are not permitted
because of safety considerations. Transport and customs arrangements for
living specimens are entirely the responsibility of the student.
Students also attending the 3D Image Processing Course, may be able to
analyze, process and display some of the 3D data collected from their
specimens


****************************************************************************=
****

=46ourth Annual

3D MAGE PROCESSING WORKSHOP

The workshop will cover 3D image processing for measurement and display.
Enrollment is limited to those attending the 3D Microscopy course.
Tuition : $850 US (lunches and snacks incl.)

WHO SHOULD ATTEND?

The course is designed for biologists working with multidimensional and
possibly multicolor microscopical data sets. Getting the data is only half
of the battle. Image data in 3, 4 or even 5 dimensions may be difficult to
store let alone analyze or display. This course is to help students
understand the hardware and software aspects of this problem and give them
the techniques they need to make the best use of their data.

The course is designed for biologists who need to make measurements on 3D
microscopical data sets and then display the results in an effective
manner. The course will be taught on SGI, Macintosh and IBM-compatible
computers. A wide variety of software designed for the 3D microscopy market
will be described, demonstrated and available for use.

Workshop Organizers

o Nick White Oxford University, UK
o Felix Margadant University of Sydney, Au

=46aculty

o Ping Chin Cheng State U. of New York, Buffalo
o Rachel Errington University of Bristol, UK
o Dan Brown Bruxton Assoc., Seattle WA
o Chris MacLean Vaytek Inc., IA
o Sigrid Myrdal Multidimensional Imaging, WA

PLAN OF INSTRUCTION
Classes will meet from 8:30-12:00 and 1:00-6:00 with lecture-demonstrations
followed immediately by hands-on laboratory sessions using a variety of
workstations. Students will "learn-by-doing" with two to a machine. Lab
handouts will describe some specific exercises to be performed on "canned"
data sets. Facilities and supervision will be available until 11:00 PM, for
students to work on their own data.
****************************************
Prof. James B. Pawley, (on Sabbatical) Ph. 61-2-9351-7548/2351
Room LG 10, Madsen Building, F-09, FAX 61-2-9351-7682
University of Sydney, NSW, 2006 Australia JBPAWLEY-at-FACSTAFF.WISC.EDU
"A scientist is not one who can answer questions but one who can question
answers." Theodore Schick Jr., Skeptical Enquirer, 21-2:39





From: Jennifer J. Mohney :      jjm8-at-lehigh.edu
Date: Fri, 3 Dec 1999 14:55:30 -0600
Subject: Job Opening Lehigh University (Correction)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nestor, I could not have the webmaster do another page, so I am sending
the message again:


The Microscopy Center at Lehigh University wishes to hire a person who
has computing and electronic skills to work with electron microscopes.
Training on the microscopes will be given if needed. The formal job
description can be found on our website at
www.lehigh.edu/~inhro/jobs.html.

We are looking for a person with general computer skills (software and
hardware familiarity with Windows and Macs) and preferably with some
electronic skills ( general maintenance and repair of power supplies and
other circuits).

The person appointed will work with two other staff members to maintain
and operate the electron microscopes of the Center; will train students
in the use of the microscopes and other equipment; will assist the
department of Materials Science and Engineering in its computer needs;
will develop remote control systems for the electron microscopes and, in
general, will handle a variety or related problems.

Lehigh University offers a generous benefit package and competitive
salary.

Send resume and letter of interest along with salary requirements and
references to Jennifer Mohney, Human Resources, 428 Brodhead Avenue,
Bethlehem, Pa 18015







From: Scott Whittaker :      sdw-at-biotech.ufl.edu
Date: Fri, 3 Dec 1999 14:53:56 -0600
Subject: Re: stain for root

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



http://www.biotech.ufl.edu/icbr/emcl/db/invisicells.html


also look at some of the other invisible sample links in the TEM section of
Tips & Tricks

http://www.biotech.ufl.edu/~emcl/tips.html

Basically use eosin B or fast green



At 12:19 PM 11/30/99 -0500, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America







From: Scott Robinson :      sjrobin-at-itg.uiuc.edu
Date: Fri, 3 Dec 1999 14:39:37 -0600
Subject: automated TEM negative processor?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi You All--

I've spoken with one or two persons about this but have not collected
specifics. Can someone out there recommend (or discourage me from
attempting to buy) an automated TEM negative processor? We'd like to be
able to somehow take a stack of exposed 3.25 x 4-in. negatives (Kodak
SO-163), pop them into a machine, and have them come out developed, dry,
and 'archival.'

Please let me know what's out there.

Thank you

Scott Robinson (sjrobin-at-uiuc.edu)
Imaging Technology Group
Beckman Institute
University of Illinois at Urbana-Champaign







From: Scott Robinson :      sjrobin-at-itg.uiuc.edu
Date: Fri, 3 Dec 1999 15:08:24 -0600
Subject: automated TEM negative processor?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi You All--

I've spoken with one or two persons about this but have not collected
specifics. Can someone out there recommend (or discourage me from
attempting to buy) an automated TEM negative processor? We'd like to be
able to somehow take a stack of exposed 3.25 x 4-in. negatives (Kodak
SO-163), pop them into a machine, and have them come out developed, dry,
and 'archival.'

Please let me know what's out there.

Thank you

Scott Robinson (sjrobin-at-uiuc.edu)
Imaging Technology Group
Beckman Institute
University of Illinois at Urbana-Champaign







From: Barbara Foster :      mme-at-map.com
Date: Sat, 04 Dec 1999 14:42:19 -0500
Subject: Fluorescence resource

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For those of you attending Cell Biology (Dec 10-15, Washington DC
Convention Center):
Dr. Manfred Hubert and Dr. Steven Ross (U. Toronto) will be available to
answer questions regarding fluorescence in the EFOS booth, #546. We have
found them a valuable resource on safety issues (ex: ozone reduction),
spectral characteristics of fluorescence systems, and photobleaching.

Best regards,
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}






From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Sat, 4 Dec 1999 10:57:56 -1000 (HST)
Subject: Need email addresses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, all

I am looking for the current email addresses of

Anne Marie Cooper
Erik Pauls
Geoff Avern
C. Michael Stanley

Thanks in advance!

Aloha,
Tina

http://www.pbrc.hawaii.edu/bemf/microangela

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************






From: Joseph Passero :      jp-at-spacelab.net
Date: Sat, 04 Dec 1999 20:07:29 -0500
Subject: LM: Looking for a Leitz Dark Field Chamber......

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Has any person in this group seen or have a "Leitz Dark Field Chamber" ?

As is described in Needham's book "The Practical Use of the Microscope" on page 293.

I have a jpg file (size is 67KB), of the figure (a drawing) from the book I can send
if you would like to see what it look's like.

I did not attach the jpg file to this posting, as some of the people on the list do not
like attachment.


Thank You

Best Regards

Joseph Passero

mailto:jp-at-spacelab.net



________________________________________________________
1stUp.com - Free the Web®
Get your free Internet access at http://www.1stUp.com





From: Gordon Couger :      gcouger-at-rfdata.net
Date: Sun, 5 Dec 1999 01:46:29 -0600
Subject: Leitz projection microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



In going through some lights I bought at the university auction I came up
with
and old Leitz projection microscope. A poor jpg is at
http://www.couger.com/images/leitz.jpg

It is a cast iron X frame with loupe like optics of about 6 & 12 power It
has a condenser,lamp house and helical focus. It is Marked Ernst Leitz
Wetlser Germany. The only other mark I can find it Germany on one lens.
The body is enameled brass. From the look I would say it was made very
early this century or late last. It stands about 9 inches tall


Does anyone have any information on this? It seems to work great.

Thanks
Gordon

Gordon Couger gcouger-at-couger.com

Stillwater, OK www.couger.com/gcouger
405 624-2855 GMT -6:00 www.couger.com/gcouger









From: Diana van Driel :      dianavd-at-eye.usyd.edu.au
Date: Mon, 6 Dec 1999 10:07:24 +1100
Subject: Printers - summary of replies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I asked about a photoquality printer that was networkable and faster
than an inkjet. Here is a summary of the replies.

****************

You might consider a Tektronix 840P solid-ink printer. We use this
printer for routine printing of reports, spectra, images, and
transparencies. The printer is much faster than the Epson 740/750 and
prints on plain paper, with surprising image quality. The cost is about
$3000. For photo-quality prints we still rely on the Epson printers.

We decided on the Tektronix after evaluating Sony and Kodak full-page
dye-sub printers and the Pictography.

*****

..our department has a number of networked
Tektronix Phaser printers at various levels of quality. They are generally
a good company, although when the ink, etc. components run out they're
pretty expensive. I think we have a Phaser 330 and 550 that are decent
intermediate printers. Cost probably in range of $1000-$2000 (just guessing
because our network administrator handles all this). I also have the Epson
Stylus Photo in my lab, and while it is quicker per print than the Phaser
550, it doesn't handle either multiple prints (} 5) or any transparencies
very well. The Phaser 550 gives better overall quality prints than the
Epson, does multiples well, and does an excellent job with transparencies.
You may want to check their web site at:
http://www.tektronix.com

*****

Alps makes a dye sub printers which are very good. We have the MD 1300. The
photo quality images are excellent, but they require the special paper and
ribbons from Alps. These printer are not very expensive ( { $400), but they
are relatively slow (5 to 10 minutes /print). The Epson Stylus Photo
printers are much faster, but have a dottier image.

*****

We use a dye-sublimation printer for images. The one nice
advantage that this type of printer offers is the ability
to make publication-ready glossy prints (on glossy paper anyway).

The resolution of the printer is 300 DPI, which is more than
enough resolution for most journals; who tend to reduce the
prints anyway.

*****

Alps makes a relatively inexpensive (~$700US) dyesub. There are some
new inkjet printers which, when used with the proper paper and new
archival inks, , make not only beautiful images but also permanent
ones. There is a review of the latest inkjets in the
Computershopper magazine (some go as high as 2400x1200 dpi) and if
you do a web search on"quad-black" you will find references to
archival inks and photographic papers.

*****

I am answering your inquiry off list. We market a new product for Epson
printers that replaces the three color ink cartridges with three black ink
cartridges. For those who need highest quality grayscale printing, this
configuration greatly widens the grayscale palette. The ink will be
archival when printed on special inkjet photo paper.

The product will only work as a plug-in to PhotoShop and drivers must be
developed for each Epson printer individually.
If you specify which Epson printer model you would use for your grayscale
printing, I'll let you know if that model is supported yet.

*****

If you are not getting any useful responses here, why not
try one of the usenet groups that deal with digital
photography and/or printers. I've found lots of help
there.

I did this and got one useful reply.....

HP, and Tektronics, both make excellent color laser printers. Tektronics is
rumored to be selling this part of their business to Xerox however. This is
not a bad thing, just thought I'd pass it along. Color lasers produce
excellent quality prints, and they are much more durable!

*****************

And what have I done?.....

I sent images to both Tektronix and QMS to print out on their laser
printers. QMS Magicolor 2 EX (2400x600dpi, laser, A4) pics were
disappointing with horizontal banding on the colour images and
unnatural colour (hot pink skin tones), though the EM's looked good.
Didn't make a good impression: I'd have expected better when they
knew there was a sale imminent. The Tektronix 840P (1200dpi, solid
ink) was the most impressive of the Tek range when viewed through a
hand lens, but from a distance the prints from the Tek 840DP, 740P
(1200dpi, A4, laser) and 780GN (1200dpi, A3, laser) were equally
good. Comparing with the same prints on coated paper from the Epson
Stylus Photo (720dpi), there was very little difference; the Epson
just had the edge. For routine printing of images, the laser printer
is quite adequate. For best quality - the Epson. As the 840 prints
can't be hot-laminated, I chose the 740. Because the 780, with A3
printing, was tempting but considerably more expensive than the 740,
I'm buying an Epson 1200 ink jet as well - don't need A3 often and
this way I also get a better photoquality printer than I already
have. It's been more than paid for by shopping around for the best
price. Because of the cost of the prints on the ink jet, not having
it networkable is an advantage in a large group!

Thanks everyone,

Diana



Diana van Driel
Department of Clinical Ophthalmology
Sydney University
GPO Box 4337
Sydney NSW
AUSTRALIA 2001

Phone 61 2 938 27276/27395
Mob 0412 165 075
Fax 61 2 938 27318





From: ard-at-ansto.gov.au (Arthur Day)
Date: Mon, 6 Dec 1999 20:03:10 +1100
Subject: Re: Printers - summary of replies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



}
} I asked about a photoquality printer that was networkable and faster
} than an inkjet. Here is a summary of the replies.
}




From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 06 Dec 99 09:06:11 -0500
Subject: Re: printer - summary of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Diana,
Thanks a lot for summarizing your responses. However, I am curious
as to your original post. Did you specifically set a price limit on the
printers? I noticed that you did not have any posts regarding the Codonics=

printers. These are quite expensive ($10-12,000) but give superb
publication quality images. The model 1660 is a thermal/dye-sub printer. =
You can
print thermal images on a special paper at the cost of about $.55 per
print. These are very high quality and rival the more expensive dye-sub
prints (~$2.00 ea). However, color must be done using the dye sublimation
method and can be with or without a protective overcoat (~$1.00 additional
cost).
The other thing I like about this printer is that it is a stand-alone
UNIX printer which permits you to adjust gamma, contrast, and color from
the printer end....so you do not have to continually adjust the original
file unless you want to go this route. You also can do various printer
formating as multiple images to a page from separt files, captions, etc. I=

tend to format pages with different sizes of images in PhotoShop but you
can also do this by programming the printer. You can also keep track of
each users printing for easy billing. This may be helpful for some networke=
d
printers. =20
I have no direct interest in this product...just a satisfied customer.

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University =20
1057 Whistler Building
West Lafayette, IN 47907-1057


Diana van Driel wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20

} Mime-Version: 1.0
} X-Sender: diana-at-mail.eye.usyd.edu.au
} Message-Id: {v04210100b470a0627a6e-at-[129.78.203.31]}
} Date: Mon, 6 Dec 1999 10:07:24 +1100
} To: MicroscopyList {microscopy-at-Sparc5.Microscopy.Com}
} From: Diana van Driel {dianavd-at-eye.usyd.edu.au}
} Subject: Printers - summary of replies
} oscopy-request-at-sparc5.microscopy.com
=20
=03=15=F2=81=B4 =14 =20
} =E8=FF =0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C=20
=0C
} =20






From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 06 Dec 99 10:58:11 -0500
Subject: Vacuum tubes responses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hi all,
I just wanted to thank everyone who responded to my post:

"We are cleaning out our shelves and want to try to sell a couple of
hundred vacuum tubes of the types used in the power cabinets of the Philips
EM-200 transmission EM. Many of the tubes have never been used but we also
have some used ones from a number of old power cabinets in storage.

We would like to find out if there is still a market value for the
tubes and where we should advertise them so they find a new owner rather
than go to the local landfill. We would like to recoup some of the cost of
the tubes if possible to help offset installation costs of some new
equipment."

I received many responses with some of the best advise to go to the web
as there are many vacuum tube venders out there. I did just that and found
out that there is indeed a thriving market for vacuum tubes...primarily
in the larger power tubes used in the old TEM power cabinets. So don't
throw them away!! Do hit the web and you may be surprised at the value of
some of these old components.

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057






From: Shotsberger-Gray, Wanda :      WandaShotsberger-Gray-at-hmhs.com
Date: Tuesday, November 30, 1999 11:19AM
Subject: stain for root

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
id xma010527; Mon, 6 Dec 99 10:14:47 -0600
Received: by SVARLEXC05 with Internet Mail Service (5.5.2448.0)
id {X6ZN316Y} ; Mon, 6 Dec 1999 10:13:48 -0600
Debby Sherman {sherman-at-btny.purdue.edu}


Debby,
Have you tried Eosin? Just add enough to color the fixative a nice bright
pink, and process as usual. Another way to do it is to add make a 1.5%
solution of cobalt chloride in absolute ethanol, and using this as a stock
solution, add 10 ml per 100 ml of absolute when processing. this colors the
tiny stuff while dehydrating. Both of these methods are common in routine
histology labs and do not interfere with immuno.
Wanda Shotsberger
Harris Methodist Hospital
Fort Worth TX
----------
} From: Debby Sherman
To: message to: MSA list
-----------------------------------------------------------------------.


Hi,
I was just given some VERY small roots to be fixed and embedded for
ICC. Fixation is not a problem but I am not sure I can keep track of these
very tiny root pieces once they are infiltrated and embedded in LR White.
I would like to stain them with something that is not soluble in ETOH
and will also not interfer with the ICC labeling.
Any suggestions?
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057






From: Wright, John D. :      jwright-at-dugway-emh3.army.mil
Date: Mon, 6 Dec 1999 09:21:24 -0700
Subject: SEM of powders

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Ladies and Gentlemen,

We are tasked with characterizing preparations of dry, powdered, milled
bacteria such as spores of Bacillus globigii and B. thuringensis using
our JEOL 6300FV. We have been using rather crude methods to transfer
some of the powdered preparations to stubs covered with double-sticky
tape. These stubs are then sputter-coated by a Denton Desk II. Our
primary concern is ensuring that we have transferred a representative
powder sample to the stub. That is, are all size ranges of particles
represented in the sample. The size range of interest is 1 to 20
microns. Size, shape, and number of particles are the parameters we
wish to measure (using, we hope, IP-plus).

Are any of you aware of standard or traditional protocols for handling
this kind of material? We are complete IPP novices; any suggestions in
properly using this software package to perform our analysis?


/John/

John D. Wright, Ph.D.
Microbiologist
West Desert Test Center
Dugway, UT 84022

(435) 831-3017





From: Nelson Conti :      conti213-at-sfsu.edu
Date: Mon, 6 Dec 1999 11:08:41 -0800
Subject: Re: EM course

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

JoAnn -
My two cents .. San Francisco State University (where I am presently
writing from, and completing my M.A. thesis) offers a course on Electron
Microscopy once a year in the Fall semester (late Aug - December). The
course number is Biology 741. Whether they'll take outside students is a
topic I cannot give an answer to, but you could investigate this by
contacting the person in charge, Gregory Antipa. His phone number is (415)
338-2951. This campus is located in San Francisco, CA near the Stonestown
Galleria and near Lake Merced (in the southern part of the city).
Nelson Conti
P.S. This campus has a world-wide web address: www.sfsu.edu
There is considerable information regarding class schedules, etc. on that
web page or links within it.







From: Lesley Weston :      lesley-at-interchange.ubc.ca
Date: Mon, 6 Dec 1999 11:47:18 -0800 (PST)
Subject: Removing epoxy from whole blocks.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone have a method for removing epoxy that doesn't involve sodium
ethoxide or methoxide? I need to do this so that I can decalcify some blocks
that unexpectedly turned out to have more heavily calcified areas than they
should have. Thanks.

Lesley Weston.











From: Tian_Huang-at-gillette.com
Date: Mon, 6 Dec 1999 14:08:39 -0500
Subject: cryo-SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, there,
Could anybody tell me where I can get cryo-SEM done? Thanks!


Tian





From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Mon, 6 Dec 1999 17:01:18 -0600
Subject: Chromium sputter coating/sample storage

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Greetings,

Does anyone have a preferred method for storage of chromium coated samples?
I'd guess that storage under vacuum or in an inert gas is the way to go, but
I'm curious as to any "pet" methods folks might have, if any.

Thanks!
Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine Bldg.
University of Missouri
Columbia, MO 65211
(573)882-8304
tindallr-at-missouri.edu
http://www.biotech.missouri.edu/emc/






From: electron microscope laboratory-UDSM :      emlab-at-udsm.ac.tz
Date: Mon, 6 Dec 1999 18:37:04 -0600
Subject: a 35mm or a CCD camera?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To: microscopy-at-sparc5.microscopy.com
} From: ard-at-ansto.gov.au (Arthur Day)


Hello all.
We have a LEO 910 TEM of which we want to install a Handy camera system,
apart from the available film plate camera.
Which one is better?
Thanks,
KIVAMBE.







From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Mon, 06 Dec 99 23:28:56 -0500
Subject: Source for Cryo SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Tian Huang wrote:
============================================
Hi, there,
Could anybody tell me where I can get cryo-SEM done? Thanks!
============================================
The laboratories of Structure Probe, Inc. have offered, since about 1985, as
part of our laboratory analytical services, fast turnaround cryo-SEM
services. We use Oxford/Hexland equipment and it is interfaced to a JEOL
Model 840 SEM. We would be happy to offer a proposal for any work that you
might want to be doing.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
============================================





From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Mon, 6 Dec 1999 19:04:57 -1000 (HST)
Subject: shelf life of resin components

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, all-

As other EM labs on Oahu have closed down in the past decade, I have
willingly scavenged their leftovers. Now I am faced with clearing out my
pack-rat space. A question arises about epoxy (mostly) resin components
that are from ca. 1992 and have remained unopened. Will any of them still
be OK? I have used old stocks of Epon812 that have been fine, but not the
anhydrides. I also have an old, unopened GMA kit. Can I still use this
stuff, or is it time to have it carted off?

Thanks in advance!

Mahalo,
Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************






From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Mon, 6 Dec 1999 19:06:55 -1000 (HST)
Subject: Have old manual for Philips 75 TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Clearing out my junk, I find I still have the complete manual for the
Philips 75 TEM. Does anyone want it? It has dates on it ranging from
1959 to 1964, and our user book goes through 1973...

Aloha,
Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************






From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Tue, 07 Dec 99 00:10:15 -0500
Subject: SEM powder preparation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

John D. Wright wrote:
==================================================
We are tasked with characterizing preparations of dry, powdered, milled
bacteria such as spores of Bacillus globigii and B. thuringensis using our
JEOL 6300FV. We have been using rather crude methods to transfer some of
the powdered preparations to stubs covered with double-sticky tape. These
stubs are then sputter-coated by a Denton Desk II. Our primary concern is
ensuring that we have transferred a representative powder sample to the stub
. That is, are all size ranges of particles represented in the sample. The
size range of interest is 1 to 20 microns. Size, shape, and number of
particles are the parameters we wish to measure (using, we hope, IP-plus).

Are any of you aware of standard or traditional protocols for handling this
kind of material? We are complete IPP novices; any suggestions in properly
using this software package to perform our analysis?
======================================================
It sounds like the SPI Tacky Dot™ Slides would stand a good chance of
working in this application. You really should be getting a representative
sampling of the particles present. For particles less than about 2-3 um you
would probably see some doublets or triplets, based on the use of the 15 um
dot size product.

More information about the Tacky Dot Slides can be found on URL
http://www.2spi.com/new/tacky.html

Disclaimer: SPI Supplies manufactures under license from DuPont the line of
Tacky Dot Slides so we would have a vested interest in seeing high volumes
used of these slides.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
============================================





From: Patton, David :      David.Patton-at-uwe.ac.uk
Date: Tue, 7 Dec 1999 09:43:07 +0000 (GMT Standard Time)
Subject: Re: shelf life of resin components

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Can't offer useful advice but I know the problem. When I
started here in 1989 I found that a previous technician
must have blown a budget surplus on resin kits. Bad move!
In those days a volatile component was supplied in
glass vials with plastic pop on lids. Suffice to say they
were all empty by the time I started, leaving me with
dozens of incomplete resin kits!

Dave



On Mon, 6 Dec 1999 19:04:57 -1000 (HST) Tina Carvalho
{tina-at-pbrc.hawaii.edu} wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi, all-
}
} As other EM labs on Oahu have closed down in the past decade, I have
} willingly scavenged their leftovers. Now I am faced with clearing out my
} pack-rat space. A question arises about epoxy (mostly) resin components
} that are from ca. 1992 and have remained unopened. Will any of them still
} be OK? I have used old stocks of Epon812 that have been fine, but not the
} anhydrides. I also have an old, unopened GMA kit. Can I still use this
} stuff, or is it time to have it carted off?
}
} Thanks in advance!
}
} Mahalo,
} Tina
}
} ****************************************************************************
} * Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
} * Biological Electron Microscope Facility * (808) 956-6251 *
} * University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
} ****************************************************************************
}
}

----------------------------------------
Patton, David
Email: David.Patton-at-uwe.ac.uk
"University of the West of England"






From: rmoretz-at-rdg.boehringer-ingelheim.com
Date: Tue, 7 Dec 1999 07:50:27 -0500
Subject: RE: shelf life of resin components

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tina:

My own experience has shown variable results in using older resins. If the
resins (including anhydrides) have not been opened, they may be ok. The
main thing to check upon opening is the color--DDSA and NMA discolor quite
badly. Another clue is during mixing of the components. Anhydrides that
are off will give a distinct orange to red/orange color in the final
mixture. However, when in doubt, throw it out (that's my motto).

Roger Moretz
Dept of Toxicology
Boehringer Ingelheim Pharmaceuticals

} -----Original Message-----
} From: Tina Carvalho [SMTP:tina-at-pbrc.hawaii.edu]
} Sent: Tuesday, December 07, 1999 12:05 AM
} To: Microscopy Listserver
} Subject: shelf life of resin components
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Hi, all-
}
} As other EM labs on Oahu have closed down in the past decade, I have
} willingly scavenged their leftovers. Now I am faced with clearing out my
} pack-rat space. A question arises about epoxy (mostly) resin components
} that are from ca. 1992 and have remained unopened. Will any of them still
} be OK? I have used old stocks of Epon812 that have been fine, but not the
} anhydrides. I also have an old, unopened GMA kit. Can I still use this
} stuff, or is it time to have it carted off?
}
} Thanks in advance!
}
} Mahalo,
} Tina
}
} **************************************************************************
} **
} * Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu
} *
} * Biological Electron Microscope Facility * (808) 956-6251
} *
} * University of Hawaii at Manoa *
} http://www.pbrc.hawaii.edu/bemf*
} **************************************************************************
} **
}





From: L. D. Marks :      ldm-at-apollo.numis.nwu.edu
Date: Tue, 7 Dec 1999 08:04:20 -0600 (CDT )
Subject: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


There are many advantages to having one general list for all
microscopies. However, there are also many disadvantages. My
primary interests are in the Physical Sciences, and I am not
interested in many of Biological topics that come up. As a
consequence the utility of this list for me has dropped a lot
as the volume of messages has increased.

I think it is time to split this list into two, one for Biology
and one for the Physical Sciences. Of course, emails can still
be sent to both lists if the topic is appropriate enough.

Comments?

++++++++++++++++++++++++++++++++++++++++++++++++
Laurence Marks
Department of Materials Science and Engineering
Northwestern University
fax: (847) 491-7820
mailto:l-marks-at-nwu.edu
http://www.numis.nwu.edu
++++++++++++++++++++++++++++++++++++++++++++++++






From: electron microscope laboratory-UDSM :      emlab-at-udsm.ac.tz
Date: Tue, 7 Dec 1999 08:13:04 -0600
Subject: a 35mm or a CCD camera?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all.
We have a LEO 910 TEM of which we want to install a Handy camera system,
apart from the available film plate camera.
Which one is better?
Thanks,
KIVAMBE.







From: John F. Mansfield :      jfmjfm-at-engin.umich.edu
Date: Tue, 7 Dec 1999 09:31:09 -0500
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree, it seems to be time.

At 8:04 AM -0600 12/7/99, L. D. Marks wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



Please note new FAX number.

John Mansfield PhD CPhys MInstP
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (734) 936-3352 FAX (734) 763-2282 Cellular Phone: (734) 358-7555
(Leaving a phone message at 936-3352 is preferable to 358-7555)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html
Location: Lat. 42=B0 16' 48" Long. 83=B0 43' 48"





From: Dr. Mark W. Lund :      lundm-at-acousb.byu.edu
Date: Tue, 07 Dec 1999 09:11:20 MST/MDT
Subject: RE: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Laurence,

Unfortunately, when lists split they often fall below
critical mass and die. I, too, get more posts that
I can't use than I can, but the signal/noise is
still much better than on the usenet.

Just my 2 cents.

best regards
mark

Mark W. Lund, PhD
VP Engineering } } Soft X-ray Web page http://www.moxtek.com { {
MOXTEK, Inc.
452 West 1260 North
Orem UT 84057 801-225-0930 FAX 801-221-1121
lundm-at-xray.byu.edu

"This is a YOUNG business...How can I tell you what
YOUR job is when I don't know what MINE is?" --Pogo





From: Chris Adams :      cadams-at-lanl.gov
Date: Tue, 7 Dec 1999 10:08:11 -0700
Subject: Fwd: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Date: Tue, 7 Dec 1999 09:46:39 -0600 (CDT )
} From: "L. D. Marks" {ldm-at-apollo.numis.nwu.edu}
} X-Sender: ldm-at-node_4990f
} To: Chris Adams {cadams-at-lanl.gov}
} Subject: Re: Time for more than one list?
} Mime-Version: 1.0
}
} Perhaps send this to the mailserver...?
}
} On Tue, 7 Dec 1999, Chris Adams wrote:
}
} } That sounds like an excellent idea to me. I for one toss 98% of the
} } messages being sent out courtesy of this list because they deal with things
} } I care little about including: biological specimen prep., job postings,
} } complaints about salaries..... I am also more intested in the physical
} } sciences being a Northwestern Materials Science Dept. alum. (MS in
} } Materials Science, 1981, Bruce Wessels).
} }
} } Chris
} }
} }
} } } ------------------------------------------------------------------------
} } } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } } -----------------------------------------------------------------------.
} } }
} } }
} } } There are many advantages to having one general list for all
} } } microscopies. However, there are also many disadvantages. My
} } } primary interests are in the Physical Sciences, and I am not
} } } interested in many of Biological topics that come up. As a
} } } consequence the utility of this list for me has dropped a lot
} } } as the volume of messages has increased.
} } }
} } } I think it is time to split this list into two, one for Biology
} } } and one for the Physical Sciences. Of course, emails can still
} } } be sent to both lists if the topic is appropriate enough.
} } }
} } } Comments?
} } }
} } } ++++++++++++++++++++++++++++++++++++++++++++++++
} } } Laurence Marks
} } } Department of Materials Science and Engineering
} } } Northwestern University
} } } fax: (847) 491-7820
} } } mailto:l-marks-at-nwu.edu
} } } http://www.numis.nwu.edu
} } } ++++++++++++++++++++++++++++++++++++++++++++++++
} }
} }
} }
} }
}
} ++++++++++++++++++++++++++++++++++++++++++++++++
} Laurence Marks
} Department of Materials Science and Engineering
} Northwestern University
} fax: (847) 491-7820
} mailto:l-marks-at-nwu.edu
} http://www.numis.nwu.edu
} ++++++++++++++++++++++++++++++++++++++++++++++++
}







From: Virginie Serin :      serin-at-cemes.fr
Date: Tue, 7 Dec 1999 18:03:59 +0100
Subject: A new EELS and XAS tool to play with

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html





{fontfamily} {param} Courier {/param} A new EELS and XAS tool to play with:



AT LAST, an EELS and XAS fine structure database exists at=20

http://www.cemes.fr/eelsdb/


The aim is to provide an extensive reference data collection.


This database is a compilation of outer-shell and inner-shell
excitation spectra from EELS and X-ray investigations, which will form
a catalog of fine structures for materials. It allows anyone to
download spectra or to submit data.


The database will be in continual development (data will be updated),
and, during the next six months, depending on your remarks, the
presentation may be modified. Please feel free to send any question or
idea to eelsdb-at-cemes.fr (or to sikora-at-lps.u-psud.fr or to
serin-at-cemes.fr).


Data can be submitted automatically via Internet. The parameters of the
spectra will be checked before incorporation of the data in the web
site. All spectra are welcome, whatever their quality (energy
resolution, signal to noise ratio,...). The justification for this is
that both high and modest energy resolution information can be useful
if these spectra exhibit changes in the fine structure.=20


*************************************************************************=
*****

Conditions of submission:

EELS:

- Raw data are for the moment preferred (no background subtraction, no
plural scattering deconvolution or resolution sharpening. Dark current
or gain variation correction is OK). Treated data are accepted when
processing permits hidden features to be revealed. Details about the
processing are required.

- Gatan-ELP binary or 2-column Ascii format

- For core edges, reduced thickness: 0.2 { { t/lambda { { 0.7.

- For outer shells (low-loss domain), reduced thickness: 0.3 { {
t/lambda { { 1 in order to offer features with optimised intensity.

- Associated low-loss data, recorded from the same area and in the same
conditions (convergence angle, mode, probe size, spectrometer aperture)
can also (and should) be submitted to offer users the possibility to
correct for thickness effects, when the single scattering profile is
needed). You must then submit the low loss data before the core-loss
data.

X-Ray:

- 2-column Ascii


Some parameters are required, some others are optional.=20

Set of EELS required parameters: Specimen name, Formula, Microscope,
Mode, Detector, Gun type, Convergence semi-angle, Objective aperture,
Incident beam energy, Collection semi-angle.

-------------------------------------------------------------------------=
------------------

{/fontfamily}






{center} -----------------------------------------------


{fontfamily} {param} New_York {/param} {bigger} {bigger} Virginie Serin

{/bigger} {/bigger} {/fontfamily} {fontfamily} {param} Geneva {/param} CEMES-CNR=
S,
29 Rue J. Marvig, BP 4347,

31 055 Toulouse Cedex 4, France

T=E9l: 33 (0)5 62 25 78 67, Fax : 33 (0)5 62 25 79 99,=20

e-mail: serin-at-cemes.fr

http://www.cemes.fr


{/fontfamily} EELS and X ray database : http://www.cemes.fr/~eelsdb/


-----------------------------------------------

{/center}






From: Barbara Foster :      mme-at-map.com
Date: Tue, 07 Dec 1999 12:10:17 -0500
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mark,

My early training came from the Royal Microscopical Society in the UK.
They routinely meet with joint sessions, bio & mat sci. I have to tell you
that each of these groups has a lot to learn from the other and would
seriously discourage splitting of this group.

An alternative: put an annotation in the header (ex: Bio: or MatS:.....).
That way, those of us who are interested in learning on a broader scale
can take advantage of the cross-pollination. Others who would like to stay
more focused can just "clump and dump" the "uninteresting" messages.

Best regards,
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************
MME is America's first national consortium providing
customized on-site workshops in all areas of
microscopy, sample preparation, and image analysis.

At 08:04 AM 12/7/99 -0600, L. D. Marks wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America






From: bozzola-at-siu.edu (John J. Bozzola)
Date: Tue, 7 Dec 1999 11:56:39 -0600
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since some of us are interested in both biological and physical sciences
postings, perhaps a better way to handle this would be to have contributors
code the subject with "B" for biological and "P" for physical. We could
activate filters to remove any unwanted mailings.

I wouldn't look forward to subscribing to yet another list.

####################################################################
John J. Bozzola, Ph.D., Director
Micro-Imaging and Analysis Center
750 Communications Drive - MC 4402
Southern Illinois University
Carbondale, IL 62901 U.S.A.
Phone: 618-453-3730
Fax: 618-453-2665
Email: bozzola-at-siu.edu
Web: http://www.siu.edu/departments/shops/cem.html
####################################################################







From: FRANK KARL :      fskarl-at-goodyear.com
Date: Tue, 7 Dec 1999 12:59:53 -0500
Subject: Re: Fwd: Re: Time for more than one whine

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I for one toss 98% of the
} } messages being sent out courtesy of this list because they deal with=
things
} } I care little about including: biological specimen prep., job postin=
gs,
} } complaints about salaries.....

Working in the rubber industry, I discard more then 99% of all the info=
rmation
listservers send me, including this one. So what! I mean it's not lik=
e I have
to scrub out each message one character at a time. I read what I find
interesting and discard the rest. My time (hope my boss doesn't see th=
is) is
not so valuable that extra 15 to 30 minutes a day can't be spend on dis=
carding
unwanted E-mail, and I often find "unrelated" information which is of v=
alue to
me.

I suspect there is some expense and effort in running and hosting a lis=
t
server, if you which to start your own, great, but let's not ask other =
people
to spend their time and effort to save us a few extra key strokes. Let=
's keep
this server as it is and if you want, you can start your own list and I=
will be
happy to subscribe to it.

Best wishes ..... Frank

=20
=





From: Ladd Research :      ladres-at-worldnet.att.net
Date: Tue, 07 Dec 1999 13:05:37 -0500
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} L. D. Marks wrote:
} }
}
} } There are many advantages to having one general list for all
} } microscopies. However, there are also many disadvantages. My
} } primary interests are in the Physical Sciences, and I am not
} } interested in many of Biological topics that come up. As a
} } consequence the utility of this list for me has dropped a lot
} } as the volume of messages has increased.
} }
} } I think it is time to split this list into two, one for Biology
} } and one for the Physical Sciences. Of course, emails can still
} } be sent to both lists if the topic is appropriate enough.
} }
} } Comments?
} }
} } ++++++++++++++++++++++++++++++++++++++++++++++++
} } Laurence Marks
I would vote to keep one general list. You can often pick up useful
information from threads you wouldn't ussually read. Plus, I don't
think you are going to find another volunteer as good as Nestor to run
another list.

Just my opinion,
JD Arnott


--

LADD RESEARCH
131 Dorset Lane
Williston, VT 05495

TEL 1-800-451-3406 (US) or 1-802-878-6711 (anywhere)
FAX 1-802-878-8074
e-mail ladres-at-worldnet.att.net
web site http://www.ladd.cc





From: William McManus :      billemac-at-biology.usu.edu
Date: Tue, 7 Dec 1999 11:21:42 -0700
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I disagree, I find it very helpful to see all of the ideas which people
have. If there is something that I am not interested in, the delete =
key is
only a keystroke away.

William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920


-----Original Message-----
} From: John F. Mansfield [mailto:jfmjfm-at-engin.umich.edu]
Sent: Tuesday, December 07, 1999 7:31 AM
To: microscopy-at-Sparc5.Microscopy.Com



I agree, it seems to be time.

At 8:04 AM -0600 12/7/99, L. D. Marks wrote:
} -----------------------------------------------------------------------=
-
} The Microscopy ListServer -- Sponsor: The Microscopy Society of =
America




From: Ann-Fook Yang (Ann-Fook Yang) :      yanga-at-em.agr.ca
Date: Tue, 07 Dec 1999 13:38:18 -0500
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I agree with Barbara.


Ann Fook Yang
EM Unit,
Eastern Cereal and Oilseed Research Centre,
Rm 2091, K.W. Neatby Bldg.,
Central Experimental Farm,
Ottawa, Ontario, Canada K1A 0C6

Phone: 613-759-1638
Fax; 613-759-1701

} } } Barbara Foster {mme-at-map.com} 12/07 12:10 PM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20

On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html=20=

-----------------------------------------------------------------------.


Mark,

My early training came from the Royal Microscopical Society in the UK.
They routinely meet with joint sessions, bio & mat sci. I have to tell =
you
that each of these groups has a lot to learn from the other and would
seriously discourage splitting of this group.

An alternative: put an annotation in the header (ex: Bio: or MatS:.....).=

That way, those of us who are interested in learning on a broader scale
can take advantage of the cross-pollination. Others who would like to stay
more focused can just "clump and dump" the "uninteresting" messages.

Best regards,
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com=20
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************
MME is America's first national consortium providing
customized on-site workshops in all areas of
microscopy, sample preparation, and image analysis.

At 08:04 AM 12/7/99 -0600, L. D. Marks wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20=









From: Tony Garratt-Reed :      tonygr-at-mit.edu
Date: Tue, 07 Dec 1999 13:44:23 -0500
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Over the time this list has been in operation, I have often enjoyed reading
about things people outside my own field have been doing.

However, the sheer volume of biological postings means that I now delete
them all wholesale. By doing this, I am negating any possible advantage of
cross-fertilization of ideas by reading the biological postings. It also
has the disadvantage that I sometimes miss postings by them being buried in
a mass of others (for example, I found Laurie's posting on this topic only
after I read John Mansfield's reply).

On the other hand, how many topics are Instrumentation rather than materials
or biological? For example, digital imaging issues are cross-discipline.
Are all such postings to be sent to both lists? Or are we going to set up
three lists? If so, should there be only three? Where will it all end?

I think there has to be an EXTREMELY compelling reason to change what is
generally regarded as a most helpful, well-organized and successful aid to
our work. I have great sympathy for Laurie's point of view. but I don't
think it fits the description of an EXTREMELY compelling reason for change.

Tony G-R




}
}
} There are many advantages to having one general list for all
} microscopies. However, there are also many disadvantages. My
} primary interests are in the Physical Sciences, and I am not
} interested in many of Biological topics that come up. As a
} consequence the utility of this list for me has dropped a lot
} as the volume of messages has increased.
}
} I think it is time to split this list into two, one for Biology
} and one for the Physical Sciences. Of course, emails can still
} be sent to both lists if the topic is appropriate enough.
}
} Comments?
}
} ++++++++++++++++++++++++++++++++++++++++++++++++
} Laurence Marks
} Department of Materials Science and Engineering
} Northwestern University
} fax: (847) 491-7820
} mailto:l-marks-at-nwu.edu
} http://www.numis.nwu.edu
} ++++++++++++++++++++++++++++++++++++++++++++++++
}
}
}

* * * * * * * * * * * * * * * * * * * * *
* Anthony J. Garratt-Reed, M.A., D.Phil.*
* SEF Team Leader *
* MIT, Room 13-1027 *
* 77 Massachusetts Avenue *
* Cambridge, MA 02139-4307 *
* USA *
* Phone: (617) 253-4622 *
* Fax: (617) 258-6478 *
* * * * * * * * * * * * * * * * * * * * *







From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Tue, 7 Dec 1999 09:00:05 -1000 (HST)
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


One list only, please! For all the reasons given today and in the past.

Aloha,
Tina

Surf's up, weather's fine. Now if I could only get out of this EM lab...

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************






From: Tobias Baskin :      BaskinT-at-missouri.edu
Date: Tue, 7 Dec 1999 13:16:32 -0600
Subject: Re: time for more than one list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
As long as the subject line is used wisely, there is little
problem. Most folks use the subject line appropriately and it is no
big deal to delete, un-read, messages on "tripod polishing" or
"Salary comparisons", or whatever. The thing is that each one of us
has a unique cross section of interests and expertise, and many would
join both lists, causing the more catholic (small c) among us to get
even more email. And those who joined just one list would reach a
lesser gathering of wisdom.

My backscattered e's,
Tobias Baskin
_ ____ ^ __ ____ Tobias I. Baskin
/ \ / / \ / \ \ University of Missouri
/ | / / \ \ \ Biological Sciences
/___/ /__ /___ \ \ \__ 109 Tucker Hall
/ / / \ \ \ Columbia, MO 65211-7400 USA
/ / / \ \ \ voice: 573-882-0173
/ /____ / \ \__/ \____ fax: 573-882-0123





From: Giles, John E Jr (FL51) :      jegiles-at-space.honeywell.com
Date: Tue, 7 Dec 1999 14:16:47 -0500
Subject: RE: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would vote for the single list also. The titles of most topics are clear
enough to differentiate Physical Sciences from Materials without any further
reading and the delete button makes quick work of sorting the things that
I'm not interested in.

Nestor does a great job running this one and there are a lot of things that
are discussed that cross the materials/physical sciences boundaries
(printers, image software, SEM care and feeding, etc.) that may be lost in a
multiple list.

I think it's kind of funny we are voting anyway. This list is Nestor's baby
and we're here because he takes the time to provide it for us. This list
was not established as a democratic entity. There's nothing in it's
"constitution" that gives anyone a "vote". His charter is not to listen to
the masses and make changes based on their wants. You abide by the rules
and if you don't like it you're free to go off and establish your own list
at anytime. I'm sure Nestor wouldn't mind a post advertising these two new
lists whenever they're established. As for me, I'll stay with this one.

John Giles
Principal Materials Engineer
Honeywell, Inc.






From: Roberto Garcia :      rgarcia-at-unity.ncsu.edu
Date: Tue, 7 Dec 1999 14:18:31 -0500
Subject: Two Lists Modest Proposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have read the posts on having two lists and I too receive many emails that
I am not interested in. However, what about general microscopy questions?
There seems to be many of them and they are usually answered by either
biologists or material scientists. I would suggest instead setting up some
rules for posting in the subject heading. That way emails can easily be
filtered through Outlook. For example BIO should begin each biological post
and MAT and GEN can begin each materials and general post. This way there is
one list and each individual can decide on the content that he wants by
filtering out the posts. What does everybody think? Of course this is all
subject to Nestors approval.
______________________________________________
Roberto Garcia
Senior Analyst, Metallography
NC State University / Analytical Instrumentation Facility
Campus Box 7531 Room 318 EGRC
1010 Main Campus Dr.
Raleigh, NC 27695-7531
(919) 515-8628
(919) 515-6965 Fax
rgarcia-at-unity.ncsu.edu
http://spm.aif.ncsu.edu/aif
____________________________________________






From: Michael Coviello :      coviello-at-mae.uta.edu
Date: Tue, 07 Dec 1999 14:00:12 -0600
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Listservers:
I agree with Mark. Some cross-discipline microscopy never hurt anyone...I
don't think.
Regards,
Mike Coviello

"Dr. Mark W. Lund" wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Dear Laurence,
}
} Unfortunately, when lists split they often fall below
} critical mass and die. I, too, get more posts that
} I can't use than I can, but the signal/noise is
} still much better than on the usenet.
}
} Just my 2 cents.
}
} best regards
} mark
}
} Mark W. Lund, PhD
} VP Engineering } } Soft X-ray Web page http://www.moxtek.com { {
} MOXTEK, Inc.
} 452 West 1260 North
} Orem UT 84057 801-225-0930 FAX 801-221-1121
} lundm-at-xray.byu.edu
}
} "This is a YOUNG business...How can I tell you what
} YOUR job is when I don't know what MINE is?" --Pogo






From: Shotsberger-Gray, Wanda :      WandaShotsberger-Gray-at-hmhs.com
Date: Monday, December 06, 1999 10:21AM
Subject: SEM of powders

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John:
To count and sort in IPP, you will first need to calibrate your specimen
from pixels to microns, then threshold to select just the items you want to
sort. After that, measure them, then sort by whatever criteria you wish to
use. It is a fairly straightforward process, once you figure out where the
buttons are for all these steps. Feel free to contact me if you think I can
help
Wanda Shotsberger
Harris Methodist Hospital
Fort Worth TX.
----------
} From: Wright, John D.
To: 'Microscopy List Server (E-mail)'
-----------------------------------------------------------------------.


Ladies and Gentlemen,

We are tasked with characterizing preparations of dry, powdered,
milled
bacteria such as spores of Bacillus globigii and B. thuringensis using
our JEOL 6300FV. We have been using rather crude methods to transfer
some of the powdered preparations to stubs covered with double-sticky
tape. These stubs are then sputter-coated by a Denton Desk II. Our
primary concern is ensuring that we have transferred a representative
powder sample to the stub. That is, are all size ranges of particles
represented in the sample. The size range of interest is 1 to 20
microns. Size, shape, and number of particles are the parameters we
wish to measure (using, we hope, IP-plus).

Are any of you aware of standard or traditional protocols for
handling
this kind of material? We are complete IPP novices; any suggestions in
properly using this software package to perform our analysis?


/John/

John D. Wright, Ph.D.
Microbiologist
West Desert Test Center
Dugway, UT 84022

(435) 831-3017





From: Keith Ryan :      KPR-at-wpo.nerc.ac.uk
Date: Tue, 07 Dec 1999 20:41:30 +0000
Subject: RE: shelf life of resin components

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Tina

I agree that unopened epoxy items should ok. We have used resins that
are VERY old. I am currently using donated Spurr (Medium) pre-weighed
mixture from June 1991. We were using some old TAAB Embedding Resin
from 1983 a few months ago - it worked fine. I found a problem with
"bulk" bottles when I uswed to weigh out mixtures from e.g. 250 ml
bottles. The anhydrides increased in viscosity as they convert by
hydrolysis to the acids. Also, opened bottles of accelerator do go
"off" after some years.

Regards from Plymouth UK

Keith Ryan











From: Gordon Couger :      gcouger-at-rfdata.net
Date: Tue, 7 Dec 1999 03:19:59 -0800
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



-From: L. D. Marks {ldm-at-apollo.numis.nwu.edu}
}
} There are many advantages to having one general list for all
} microscopies. However, there are also many disadvantages. My
} primary interests are in the Physical Sciences, and I am not
} interested in many of Biological topics that come up. As a
} consequence the utility of this list for me has dropped a lot
} as the volume of messages has increased.
}
} I think it is time to split this list into two, one for Biology
} and one for the Physical Sciences. Of course, emails can still
} be sent to both lists if the topic is appropriate enough.
}

My interest are in LM and even less info is on this list about this.
But I am very strongly for one list. It is a small list that generates
less mail than I get on get rich quick spam every day. You don't
have to read them all.

The advantage to one list is the interdisplanary answeres to questions
And this is the best list I have found for this.

Please do FIX it is isn't broken.

Gordon
Gordon Couger
624 Cheyenne
Stillwater, OK 74075
405 624 2855 GMT - 6:00







From: SEMTRADER-at-aol.com
Date: Tue, 7 Dec 1999 16:33:05 EST
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I prefer one list also.


Michael Augustyn





In a message dated 12/7/99 2:45:44 PM Eastern Standard Time,
jegiles-at-space.honeywell.com writes:


{ { Subj: RE: Time for more than one list?
Date: 12/7/99 2:45:44 PM Eastern Standard Time
From: jegiles-at-space.honeywell.com (Giles, John E Jr (FL51))
To: Microscopy-at-Sparc5.Microscopy.Com (Microscopy List)

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America
To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
-----------------------------------------------------------------------.


I would vote for the single list also. The titles of most topics are clear
enough to differentiate Physical Sciences from Materials without any further
reading and the delete button makes quick work of sorting the things that
I'm not interested in.

Nestor does a great job running this one and there are a lot of things that
are discussed that cross the materials/physical sciences boundaries
(printers, image software, SEM care and feeding, etc.) that may be lost in a
multiple list.

I think it's kind of funny we are voting anyway. This list is Nestor's baby
and we're here because he takes the time to provide it for us. This list
was not established as a democratic entity. There's nothing in it's
"constitution" that gives anyone a "vote". His charter is not to listen to
the masses and make changes based on their wants. You abide by the rules
and if you don't like it you're free to go off and establish your own list
at anytime. I'm sure Nestor wouldn't mind a post advertising these two new
lists whenever they're established. As for me, I'll stay with this one.

John Giles
Principal Materials Engineer
Honeywell, Inc.




----------------------- Headers --------------------------------
Return-Path: {Microscopy-request-at-sparc5.microscopy.com}
Received: from rly-yd02.mx.aol.com (rly-yd02.mail.aol.com [172.18.150.2])
by air-yd01.mail.aol.com (vx) with ESMTP; Tue, 07 Dec 1999 14:45:44 -0500
Received: from Sparc5.Microscopy.Com ([206.69.208.10]) by
rly-yd02.mx.aol.com (v66.4) with ESMTP; Tue, 07 Dec 1999 14:45:31 1900
Received: (from daemon-at-localhost) by Sparc5.Microscopy.Com (8.6.11/8.6.11)
id NAA08076 for dist-Microscopy; Tue, 7 Dec 1999 13:18:31 -0600
Received: from no_more_spam.com (Sparc5 [206.69.208.10]) by
Sparc5.Microscopy.Com (8.6.11/8.6.11) with SMTP id NAA08065 for
"MicroscopyFilteredEmail-at-msa.microscopy.com"; Tue, 7 Dec 1999 13:17:59 -0600
Received: from fl51m03.space.honeywell.com (fl51m03.space.honeywell.com
[130.181.6.220]) by Sparc5.Microscopy.Com (8.6.11/8.6.11) with ESMTP id
NAA08057 for {Microscopy-at-MSA.Microscopy.com} ; Tue, 7 Dec 1999 13:17:39 -0600
Received: by fl51m03.space.honeywell.com with Internet Mail Service
(5.5.2650.21)
id {YG1J3HT4} ; Tue, 7 Dec 1999 14:18:09 -0500
Message-ID:
{BA5F9C6AA352D311AD4E009027855BCD01EF7DAC-at-fl51m01.space.honeywell.com}
From: "Giles, John E Jr (FL51)" {jegiles-at-space.honeywell.com}
To: Microscopy List {Microscopy-at-Sparc5.Microscopy.Com}
Subject: RE: Time for more than one list?
Date: Tue, 7 Dec 1999 14:16:47 -0500
MIME-Version: 1.0
X-Mailer: Internet Mail Service (5.5.2650.21)
Content-Type: text/plain;
charset="iso-8859-1"
Errors-to: Microscopy-request-at-sparc5.microscopy.com

} }





From: Sara Miller :      saram-at-duke.edu
Date: Tue, 7 Dec 1999 16:47:40 -0500 (EST)
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Amen to the two Johns. I am mostly biological, but get some tips from
the physical notes. If I'm not interested in the topic, I just hit the
delete key. I'm sure I wouldn't take the time to log on to a different
site for the physical comments; thus, I'd miss some useful stuff.

This is a good time to remind users to be concise and clear in the
subject line. A "B" or "P" is not a bad idea either.

Sara

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735






From: Joseph Passero :      jp-at-spacelab.net
Date: Tue, 07 Dec 1999 16:59:43 -0500
Subject: LM: Looking for Leitz Retical.....

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I am looking for a Leitz retical, OM 1 (18.5 to 19mm diameter) for use with the old
orthomat camera.

Thank you

Best Regards

mailto:jp-at-spacelab.net



________________________________________________________
1stUp.com - Free the Web®
Get your free Internet access at http://www.1stUp.com





From: =shAf= :      mshaf-at-darkwing.uoregon.edu
Date: Tue, 7 Dec 1999 14:23:16 -0800
Subject: RE: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


John writes ...

} ---
}
}
} I agree, it seems to be time.
}
} At 8:04 AM -0600 12/7/99, L. D. Marks wrote:
}
} ...
}
} I think it is time to split this list into two, one for Biology
} and one for the Physical Sciences. Of course, emails can still
} be sent to both lists if the topic is appropriate enough.
}
} Comments?


So far today ... these 2 messages amount to half my "microscopy list"
e-mail. If it were a busier list, I might agree ... but personally, I
don't have a problem with recognizing the posts I'm not interested in
and deleting them. I believe there is too much on this list regarding
general microscopy, as applied to all sciences, that I'd be afraid to
miss if I joined a "materials' list".

Which is not to say ... some of you could do a better job with
focussing the subject line ... some mails you have to open to realize
what's inside.

cheerios, shAf

{} /\ {\/} /\ {\/} /\ {\/} /\ cogito, ergo zZOooOM /\ {\/} /\ {\/} /\ {\/} /\ {}
Michael Shaffer, R.A. - ICQ 210524
Geological Science's Electron Probe Facility - University of Oregon
mshaf-at-darkwing.uoregon.edu - http://darkwing.uoregon.edu/~mshaf/








From: Michael Bode :      mb-at-soft-imaging.com
Date: Tue, 7 Dec 1999 15:13:48 -0700
Subject: RE: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree with Barbara. Although I just scan many of the messages that
request specific information about something I know nothing about, I
personally enjoy reading about new stuff and learning from it. I believe
shattering this list into 2 or more sublists is counterproductive.

I have set up my email application so that every mail with
"www.msa.microscopy.com" in the text gets automatically transferred into
a separate folder. When I have a few minutes (not often these days!!), I
sift through the messages, delete the ones that deal with subjects that
I don't track and move the others into a "keep" folder. It's really not
a lot of work and I know most of the time by the subject line what I
want to keep or throw out.

Just my humble opinion.

Michael

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================



} ----------
} From: Barbara Foster[SMTP:MME-at-MAP.COM]
} Sent: Tuesday, December 07, 1999 10:10:17 AM
} To: L. D. Marks; Microscopy List
} Subject: Re: Time for more than one list?
} Auto forwarded by a Rule
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Mark,

My early training came from the Royal Microscopical Society in the UK.
They routinely meet with joint sessions, bio & mat sci. I have to tell
you
that each of these groups has a lot to learn from the other and would
seriously discourage splitting of this group.

An alternative: put an annotation in the header (ex: Bio: or
MatS:.....).
That way, those of us who are interested in learning on a broader scale
can take advantage of the cross-pollination. Others who would like to
stay
more focused can just "clump and dump" the "uninteresting" messages.

Best regards,
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************
MME is America's first national consortium providing
customized on-site workshops in all areas of
microscopy, sample preparation, and image analysis.

At 08:04 AM 12/7/99 -0600, L. D. Marks wrote:
} -----------------------------------------------------------------------
-
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Keith Ryan :      KPR-at-wpo.nerc.ac.uk
Date: Tue, 07 Dec 1999 22:50:24 +0000
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm happy to continue hitting the "Delete" button - but sometimes its
good to look into other fields - you never know what you'll find.

Keith Ryan
Marine Biological Association
Plymouth UK





From: Diana van Driel :      dianavd-at-eye.usyd.edu.au
Date: Wed, 8 Dec 1999 10:04:15 +1100
Subject: RE: Printers - summary of replies

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Debby,

I wanted a moderately fast networkable printer and specifically
mentioned lasers, photoquality and not too expensive. Prices in Aus
are higher than in the US, so what is expensive to you is very
expensive for us. The laser printers have the advantage of being able
to do routine Departmental printing.

Diana



Diana van Driel
Department of Clinical Ophthalmology
Sydney University
GPO Box 4337
Sydney NSW
AUSTRALIA 2001

Phone 61 2 938 27276/27395
Mob 0412 165 075
Fax 61 2 938 27318





From: Palatsides, Manuela :      m.palatsides-at-pmci.unimelb.edu.au
Date: Wed, 8 Dec 1999 10:17:33 +1100
Subject: EM: background on slot grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are having problems with terrible background when immunogold labelling
sections on slot grids using either formvar or butvar as support films..
Using the exact same procedure on 300 hexagonal grids with the same support
film background is not a problem.

The background in on the resin, all the cells and support film.

Has anyone else experienced this problem using slot grids?

Manuela Palatsides
Electron Microscopy
Peter MacCallum Cancer Institute
Locked Bag#1
A'Beckett Street
Melbourne 3000

Telephone: 03 96561244
Fax: 03 96561411
Email: m.palatsides-at-pmci.unimelb.edu.au







From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Tue, 7 Dec 1999 17:59:53 -0600
Subject: film dryers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All,
We are in the market for a small film drying cabinet for our EM film. We
don't need, or have space for one of thoses locker-sized jobs they sell in
photo suplly catalogs. Any ideas on where to look? Or better still, is
anyone looking to get rid of one?

Thanks,
Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Confocal Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175







From: Nestor J. Zaluzec :      zaluzec-at-aaem.amc.anl.gov
Date: Tue, 7 Dec 1999 18:09:24 -0600
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



This comment comes up at least once or twice/year generally in private messages
to me, but occasionally in the public forum.

Besides wearing out your delete button what are the disadvantages
of allowing the interaction of all microscopists, regardless of their
background?

The free and mutual exchange of ideas and input ESPECIALLY from
those of different disiciplines IMHO far outweighs the minor annoyance of
deleting
messages. The way our understanding expands is by hearing and listening
to what other in different fields say, and from that dialog coming up with
new knowledge. The long term benefits of that are potentially too great to
ignore. We do not exist in isolation and cannot learn from others if we create
isolation. Clearly mixing commentary on a completely unrelated subjects (say
automobile repair with microscopy) has very little potential for creating
new understanding, but mixing light, electron, probe, and x-ray
microscopy/microanalysis between life and physical sciences is a much
different issue.

The appropriate use of the SUBJECT field is important in reducing/helping
others decide on what to read or not read and I would (as always) encourage
better use.
But that was discussed when the server started back in 1993 !! Has it
really been that long...?? Yep I guess so. In any case it's been alot of
Mbytes of texts that's for sure.

So in short, I would disagree with those who think the list should be divided
up. I think too much will be lost from cross-fertilization of ideas and
concepts. The delete button works very well for me and I probably get alot more
messages than most of you on the server.

But, that is my personal opinion. I'm interested
in the comments of others on the list and will sit back and listen
for a bit to see how the discussion develops within our "little group".

Nestor
Your Friendly Neighborhood SysOp.





==================================================================
Nestor J. Zaluzec
Materials Science Division
Building 212
Argonne National Lab
9700 S. Cass Ave
Argonne, Illinois 60439 USA
Tel: 630-252-7901, Fax: 630-252-4798
Email: Zaluzec-at-aaem.amc.anl.gov
==================================================================
TPMLab: http://tpm.amc.anl.gov
MMSite: http://www.amc.anl.gov
==================================================================

The box said "This program requires Win 95/98/NT or better..." so I bought
a G3 Mac

==================================================================







From: Augusto_A_Morrone-at-notes.seagate.com
Date: Tue, 7 Dec 1999 18:09:38 -0600
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I scan messages by subject, thus reducing to a minimum the number of messages
that I start to read. The first few lines of the actual message are another
effective filter. This works well for me.

Two lists will not split subjects but rather people (subscribers) into two
groups. Since some subjects are of interest to both physical and biological
subscribers (ex: instrumentation, printers, scanners, image manipulation,
conferences covering both aspects of microscopy, cameras, and some microscopy
techniques), such a division may not turn out to be a good thing. Would MSA
eventually split into two societies too?

Just writing clear subject lines in every message, and using some discretion in
the number of postings as well as in the length of discussions - like this one -
will keep the listserver effective for everyone.

Augusto







From: Huggins, Bradley J :      HUGGINBJ-at-BP.com
Date: Tue, 7 Dec 1999 19:23:04 -0500
Subject: RE: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


True, the list is a very busy place. This can be a bit of a burden, but it
is usually easy enough to decide if the topic is of interest to you from the
subject, or at least, the title or first line of the text. But I think that
this is one of its greatest advantages. We all can share our problems and
experience right, here right now. I could see some categorizing of
messages. For example, we could require the use of one or more keywords in
the subject such as: biological/materials/physical science, or microscope
instrumentation/theory. But, if we do separate the list, we will loose a
tremendous strength, that is the very depth of the support that is available
on this list.

Brad Huggins
BPAmoco, Naperville, IL
Microscopy & Microanalysis Lab



} ----------
} From: L. D. Marks[SMTP:ldm-at-apollo.numis.nwu.edu]
} Sent: Tuesday, December 07, 1999 8:04 AM
} To: Microscopy List
} Subject: Time for more than one list?
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} There are many advantages to having one general list for all
} microscopies. However, there are also many disadvantages. My
} primary interests are in the Physical Sciences, and I am not
} interested in many of Biological topics that come up. As a
} consequence the utility of this list for me has dropped a lot
} as the volume of messages has increased.
}
} I think it is time to split this list into two, one for Biology
} and one for the Physical Sciences. Of course, emails can still
} be sent to both lists if the topic is appropriate enough.
}
} Comments?
}
} ++++++++++++++++++++++++++++++++++++++++++++++++
} Laurence Marks
} Department of Materials Science and Engineering
} Northwestern University
} fax: (847) 491-7820
} mailto:l-marks-at-nwu.edu
} http://www.numis.nwu.edu
} ++++++++++++++++++++++++++++++++++++++++++++++++
}
}





From: Linda Rangell :      lkc-at-gene.com
Date: Tue, 07 Dec 1999 16:25:46 -0800
Subject: TEM on viruses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know of a contract lab in the SF Bay Area that will do TEM
of virus particles? I'm not sure if the person I'm asking for would
like negative staining or thin sections.

Thanks for any information,
Linda Rangell






From: Laurie Wallin :      lwallin-at-ucsd.edu
Date: Tue, 7 Dec 1999 16:58:09 -0800
Subject: new list ideas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree with Barbara Foster's comment. I am working in a branch of biology
and though I do not understand the whole of the physical microscopy
workings, I enjoy hearing about interesting topics no matter what field. I
like the idea to attach a heading to the front of subject lines that would
designate messages as Bio, versus Mat. Sci.

Sincerely,
Laurie Wallin
UCSD Department of Anesthesiology
9500 Gilman Drive, 0629, La Jolla, CA 92093
(858) 822-3271





From: L. D. Marks :      ldm-at-apollo.numis.nwu.edu
Date: Tue, 7 Dec 1999 18:59:09 -0600 (CDT )
Subject: Summary: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have been counting responses either way, some of
which have not gone to the listserver. At the
moment it is about 3:1 against a split (sigh), with a
strong sentiment for better subject lines.

++++++++++++++++++++++++++++++++++++++++++++++++
Laurence Marks
Department of Materials Science and Engineering
Northwestern University
fax: (847) 491-7820
mailto:l-marks-at-nwu.edu
http://www.numis.nwu.edu
++++++++++++++++++++++++++++++++++++++++++++++++






From: COURYHOUSE-at-aol.com
Date: Tue, 7 Dec 1999 20:22:25 EST
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I vote for one list ... there is not to many messages really, I am on some
list servers that generate 60 to 200 a day, now that is excessive!

I for one, am interested in both bio and material applications of both light
and em
so one list is fine by me.

Ed Sharpe archivist for SMECC





From: p00bare :      p00bare-at-pdq.net
Date: Tue, 07 Dec 1999 22:04:30 -0600
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I'm a geologist/mineralogist myself, but I find many of the biological
postings useful, especially those on sample prep. Certainly experience
with printers or archiving is a general topic, as are details on
particular microscopes. Me...I'll take the whole enchilada; easy to
discard what does not sound useful. Dave Pevear, Houston

"L. D. Marks" wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} There are many advantages to having one general list for all
} microscopies. However, there are also many disadvantages. My
} primary interests are in the Physical Sciences, and I am not
} interested in many of Biological topics that come up. As a
} consequence the utility of this list for me has dropped a lot
} as the volume of messages has increased.
}
} I think it is time to split this list into two, one for Biology
} and one for the Physical Sciences. Of course, emails can still
} be sent to both lists if the topic is appropriate enough.
}
} Comments?
}
} ++++++++++++++++++++++++++++++++++++++++++++++++
} Laurence Marks
} Department of Materials Science and Engineering
} Northwestern University
} fax: (847) 491-7820
} mailto:l-marks-at-nwu.edu
} http://www.numis.nwu.edu
} ++++++++++++++++++++++++++++++++++++++++++++++++






From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Tue, 07 Dec 1999 20:26:08 -0800
Subject: RE: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think some abbreviations to indicate particular microscopy area in the
beginning of the "Subject" line will be a plus. It does not hurt the idea
of solid ListServer but give us some flexibility to set up our E. mail
programs to filter and organize information better. Expanding this idea I
would suggest Nestor may establish some "rules" for "Subject" line. For
example, abbreviation should have a 4 letters starting with capital and
ending with ":". Something like "Bio:"; "Mat:", "Gen:". Capital and ":"
or something like that may help to tune up filters more precisely. I
think, there are more than 3 categories may be established. If for some
reason people do not accept such rules - it did not interrupt ListServer
functionality.


_________________________________

Sergey Ryazantsev Ph. D.
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu
http://www.bol.ucla.edu/~sryazant







From: Walck. Scott D. :      walck-at-ppg.com
Date: Wed, 8 Dec 1999 00:25:43 -0500
Subject: RE: film dryers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You can make a cheap one from Plexiglas and a hair dryer. Make it to fit
with your film processing rack, with the rack off the bottom (two rods will
work or you can make a plenum), put the dryer in the bottom in a hole made
to fit with a collar, and you need a top with holes. I was going to have
one made here, but the machinist that I put on it made the same thing from
stainless steel. More expensive, but it works very well. It has the
footprint of about 8"x24". (The hair dryer is along the long axis.)

If you have the money, there is a company called California Stainless that
makes some of the ones that are sold in the EM catalogs. You can buy direct
from them. Sorry, I don't have the address.
-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)


"The opinions expressed are those of Scott D. Walck and not of PPG
Industries, Inc. nor of any PPG-associated companies."
--


} -----Original Message-----
} From: Leona Cohen-Gould [mailto:lcgould-at-mail.med.cornell.edu]
} Sent: Tuesday, December 07, 1999 7:00 PM
} To: Microscopy-at-Sparc5.Microscopy.Com
} Subject: film dryers
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.} html
}
}
} --------------------------------------------------------------
} ---------.
}
}
} Hi All,
} We are in the market for a small film drying cabinet for our
} EM film. We
} don't need, or have space for one of thoses locker-sized jobs
} they sell in
} photo suplly catalogs. Any ideas on where to look? Or
} better still, is
} anyone looking to get rid of one?
}
} Thanks,
} Lee
}
} Leona Cohen-Gould, M.S.
} Sr. Staff Associate
} Director, Electron Microscopy Core Facility
} Manager, Confocal Microscopy Core Facility
} Joan & Sanford I. Weill Medical College
} of Cornell University
} voice (212)746-6146
} fax (212)746-8175
}
}
}





From: r.cross-at-ru.ac.za
Date: Wed, 8 Dec 1999 09:27:02 +0200
Subject: one or two lists?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings to everyone with all microscopy interests!

I believe most strongly that there should remain only one list. I am
sure that had there been two or more lists I would have missed
some important pieces of information. While the amount of traffic
on the single list remains manageable (15 - 20 messages a day is
easy to handle in a few minutes) lets keep it this way.

Many thanks, Nestor, for your good work.

Regards



Robin H Cross
Director : EM Unit, Rhodes University, Grahamstown, South Africa
R.Cross-at-ru.ac.za
tel: +27 46 603 8168 - fax: +27 46 622 4377
http://www.ru.ac.za/emu/index.htm

** remember that ICEM-15 in 2002 is in Durban, South Africa **





From: Gordon Couger :      gcouger-at-rfdata.net
Date: Wed, 8 Dec 1999 05:44:32 -0600
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Leave it the way it is. Better subjects will save us all time.
But don't break up the best list for finding a solution to
a problem I have found.

HIP Hippo Hip Hooray NESTER

It should be possible to write software to block the mail
you don't want sent a the mail server.

It would also be possible to send it to a second mailer and
strip out what you don't want or use the filter in your mail
client to filter out the unwanted postings.

I might be talked into doing it but I real think that if numbers
are a problem go to a digest.

As far as this being a busy list is a joke. While it has the highest signal
to noise of any group I have been on it is very low volume list.

I read 50 to 70 messages out ot 250 and answer or originate about 25 a day.
30% of this is joke and other personal stuff. Since I am retired I don't
mind a
bit.


Gordon







From: Ronald Smith :      rsmith-at-julian.uwo.ca
Date: Wed, 08 Dec 1999 08:11:15 -0500
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree, one list only please. Cross subject information is a means of
broadening our appreciation of the many applications of microscopy. in my
lab where we service not only biological materials primarily but also soil
and mineral from our geology department the single list is most useful.

Ronald Smith,
Plant Sciences Department,
University of Western Ontario,

Gordon Couger wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} -From: L. D. Marks {ldm-at-apollo.numis.nwu.edu}
} }
} } There are many advantages to having one general list for all
} } microscopies. However, there are also many disadvantages. My
} } primary interests are in the Physical Sciences, and I am not
} } interested in many of Biological topics that come up. As a
} } consequence the utility of this list for me has dropped a lot
} } as the volume of messages has increased.
} }
} } I think it is time to split this list into two, one for Biology
} } and one for the Physical Sciences. Of course, emails can still
} } be sent to both lists if the topic is appropriate enough.
} }
}
} My interest are in LM and even less info is on this list about this.
} But I am very strongly for one list. It is a small list that generates
} less mail than I get on get rich quick spam every day. You don't
} have to read them all.
}
} The advantage to one list is the interdisplanary answeres to questions
} And this is the best list I have found for this.
}
} Please do FIX it is isn't broken.
}
} Gordon
} Gordon Couger
} 624 Cheyenne
} Stillwater, OK 74075
} 405 624 2855 GMT - 6:00








From: Gillmeister, Russ :      RGillmeister-at-sdms.usa.xerox.com
Date: Wed, 8 Dec 1999 08:37:27 -0500
Subject: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Larry, I agree, although I'm not the one doing the work maintaining this
list.
Thanks Nestor! Russ Gillmeister, Xerox

-----Original Message-----
} From: L. D. Marks [mailto:ldm-at-apollo.numis.nwu.edu]
Sent: Tuesday, December 07, 1999 9:04 AM
To: Microscopy List


There are many advantages to having one general list for all
microscopies. However, there are also many disadvantages. My
primary interests are in the Physical Sciences, and I am not
interested in many of Biological topics that come up. As a
consequence the utility of this list for me has dropped a lot
as the volume of messages has increased.

I think it is time to split this list into two, one for Biology
and one for the Physical Sciences. Of course, emails can still
be sent to both lists if the topic is appropriate enough.

Comments?

++++++++++++++++++++++++++++++++++++++++++++++++
Laurence Marks
Department of Materials Science and Engineering
Northwestern University
fax: (847) 491-7820
mailto:l-marks-at-nwu.edu
http://www.numis.nwu.edu
++++++++++++++++++++++++++++++++++++++++++++++++





From: Michael BUCKER :      MBUCKER-at-dgs.state.va.us
Date: Wed, 08 Dec 1999 08:59:45 -0500
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Regarding the proposed list split:
I myself am involved in both material and biological microscopy. My work =
involves identifying sources of animal, vegetable, and mineral "unknowns" =
via various light microscopy methods of examination. In my opinion, it =
would do well for my purposes to split the list into "TEM" and "Others" =
rather than "materials" and "biological". I say this because it seems =
that 90% of the emails on this list server are related to TEM applications =
which are generally all Greek to me and serve no useful purpose in my =
field. Still, if this server remains "as is", my 5 minutes a day deleting =
obvious TEM related messages is no big problem. As others have stated, =
keeping the Subject narrow and Specific must be the goal if only one site =
is to remain available.
My 3 cents (inflation adjusted)
Mike Bucker
Microscopy Principal
Consolidated Labs of Virginia
=20

} } } "Nestor J. Zaluzec" {zaluzec-at-aaem.amc.anl.gov} 12/07 7:09 PM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20

On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html=20=

-----------------------------------------------------------------------.



This comment comes up at least once or twice/year generally in private =
messages
to me, but occasionally in the public forum.

Besides wearing out your delete button what are the disadvantages
of allowing the interaction of all microscopists, regardless of their
background?

The free and mutual exchange of ideas and input ESPECIALLY from
those of different disiciplines IMHO far outweighs the minor annoyance of
deleting
messages. The way our understanding expands is by hearing and listening
to what other in different fields say, and from that dialog coming up =
with
new knowledge. The long term benefits of that are potentially too great to
ignore. We do not exist in isolation and cannot learn from others if we =
create
isolation. Clearly mixing commentary on a completely unrelated subjects =
(say
automobile repair with microscopy) has very little potential for creating
new understanding, but mixing light, electron, probe, and x-ray
microscopy/microanalysis between life and physical sciences is a much
different issue.

The appropriate use of the SUBJECT field is important in reducing/helping
others decide on what to read or not read and I would (as always) =
encourage
better use.
But that was discussed when the server started back in 1993 !! Has it
really been that long...?? Yep I guess so. In any case it's been alot of
Mbytes of texts that's for sure.

So in short, I would disagree with those who think the list should be =
divided
up. I think too much will be lost from cross-fertilization of ideas and
concepts. The delete button works very well for me and I probably get alot =
more
messages than most of you on the server.

But, that is my personal opinion. I'm interested
in the comments of others on the list and will sit back and listen
for a bit to see how the discussion develops within our "little group".

Nestor
Your Friendly Neighborhood SysOp.





=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D
Nestor J. Zaluzec
Materials Science Division
Building 212
Argonne National Lab
9700 S. Cass Ave
Argonne, Illinois 60439 USA
Tel: 630-252-7901, Fax: 630-252-4798
Email: Zaluzec-at-aaem.amc.anl.gov=20
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D
TPMLab: http://tpm.amc.anl.gov=20
MMSite: http://www.amc.anl.gov=20
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D

The box said "This program requires Win 95/98/NT or better..." so I =
bought
a G3 Mac

=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=
=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D=3D









From: Kristian Ukkonen :      kukkonen-at-cc.hut.fi
Date: Wed, 08 Dec 1999 16:10:10 +0200
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



People,

Perhaps you should read from the list archives the previous
discussions about this very same subject.. It appears that
this same discussion surfaces every few years, and ends to
the same conclusion..

Besides that, the "idea" of LM: EM: etc. prefixes in subject
lines is VERY old, and in fact what people SHOULD be doing
all the time. RTFFAQ. :) Read the fine FAQ. From FAQ:
".. Also preface your description by the conventional abbreviation
of the type of microscopy you are interested such as LM, IRM, XRM,
TEM, SEM, AFM, STM, uProbe......"

Just to add something useful to this message, I recently
copied all old list archive files into one file (about
50 megabytes), and now if I need to check something, I
just do a search to the single file, and find all old
discussions.. Very useful. Unix people would use less and
/ command, PC people list and \ command..

Quick search revealed that this discussion has been on:
October 93, May 95, Dec 99..

Best regards,

Kristian Ukkonen.





From: Wayne England :      wengland-at-ortech.on.ca
Date: Wed, 8 Dec 1999 09:33:13 -0500
Subject: eXL monitor - thanks for the input

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This thread is a fairly hardy perennial and will be familiar to most members
who have been on the list awhile, so just to add my 0.0126 Euro worth.

The concept of specific subjects lines for the list is fine.

Indeed we are asked to do this by Nestor in his original welcome text.

To quote from the saved message (Well, some of us saved it!) :-
======================================================
As a courtesy to the readers of this list please indicate in the subject
line of your message a reasonable descriptive title of your comment/inquiry.
Also preface your description by the conventional abbreviation of the type
of microscopy you are interested such as LM, IRM, XRM, TEM, SEM, AFM, STM,
uProbe...... For example, if you are interested in optical microscopy and
have a question about staining then a Title/Subject line for your message
might be


Thanks to all who responded to the crisis of repairing or replacing our eXL
monitor. We were eventually able to locate a local TV repair shop that had
it corrected in a couple of days at a relatively meager price of $300.00.
It was nice to have some potential options however, should this attempt have
failed.

While I'm here. I agree with Nestor's assessment of the listserver
objectives. I would much rather filter out the non-relevant information
than miss a treasured tip on managing a difficult process or sample prep.
The boundaries of science are not that distinct. Remember, everything in
life is relative!! My two cents.


============================================
Wayne England
Manager, Physical Characterization
Bodycote ORTECH Inc.
2395 Speakman Drive, Mississauga, ON, L5K1B3
wengland-at-ortech.on.ca WEB: www.bodycote.com
905-822-4111 Ext.555 FAX:905-823-1446
============================================






From: Karen S Pawlowski :      kna101-at-utdallas.edu
Date: Wed, 8 Dec 1999 08:46:00 -0600 (CST)
Subject: Re: Removing epoxy from whole blocks.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lesley,

You may not need to remove the expoy to remove the calcium. Depending on
how big the block is and how much of it needs to be cut, you may be able
to emerse it in EDTA for an extended amount of time to remove enough
calcium for sectioning. I have never done this myself, but have seen
several people do it over the years. Maybe someone out there with more
direct experience will respond.....

Karen


On Mon, 6 Dec 1999, Lesley Weston wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Does anyone have a method for removing epoxy that doesn't involve sodium
} ethoxide or methoxide? I need to do this so that I can decalcify some blocks
} that unexpectedly turned out to have more heavily calcified areas than they
} should have. Thanks.
}
} Lesley Weston.
}
}
}
}
}
}
}
}






From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 08 Dec 99 09:52:00 -0500
Subject: Bio-summary-stain for root

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Thanks to all who sent suggestions for staining root so it was easier to
see when embedded in LR White. Most called for a stain added when
tissue was in 100% ETOH. Since I had very little material and did not want=
to
loose it during dehydration, I decided to try toluidine blue after
fixation and prior to ETOH to see if it would tolerate the dehydration, etc=
. I
was delighted with the end result. A couple of drops of a 1% solution
added to the post fix buffer gave sufficient staining after only about 5-10=

minutes to carry all the way through the embedding process. I do not know
if there was any effect on immuno-reactivity yet....let's hope not.

In any case, additional suggestions are summarized below. A couple,
esp. the method of enrobbing small specimens with formvar (T. Baskin) was
especially intriging. I certainly will try that method. This list is an
endless stream of good ideas!!
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University =20
1057 Whistler Building
West Lafayette, IN 47907-1057

Phluoroglucinol. Excellent stain for lignified structures. Could also
use
Graff's "C" stain which is a standard for wood and paper fiber
structures.
Lou Solebello
-------------
Re: stain for rootGreetings,=A0=A0=A0=A0=A0=A0=A0 Debby Sherman asked about=
handling
small bits of root. I make my living handling these snippets so perhaps I
can help.
=A0=A0=A0=A0=A0=A0=A0 As far as stains go, we have used two. We have made a=
saturated
solution of Fast Green in ethanol. We make an 8% solution of fast green in
100% ethanol and then add 1-2 =B5l to each half ml of volume with sample. I=

have also used a 0.15 % aqueous solution of basic fuscin. In this case i
collect the samples in that prior to dehydration, and they stay nicely
purple throughout.
=A0=A0=A0=A0=A0=A0=A0 The other thing that I do to greatly simplify my life=
is to put
each root on a wire loop, encased in Formvar. Here is an excerpt from a lab=

protocol I have:
=A0=A0=A0=A0=A0=A0=A0We get best results with a Formvar loop method. In thi=
s method, a
loop of copper wire (36 gauge) is made and flattened between two flat
pieces of steel. Then small rectangles of 0.25% Formvar in ethylene chlorid=
e
are floated on water and the loop plunged into the middle of the rectangle
so that a film of Formvar surrounds the wire loop. A number of loops are
made in advance (can be days ahead). A root is then placed on the Formvar
surface, the excess cut away with a razor and then this assembly is coated
with another Formvar layer, in the same was as above, thus encasing the
root between Formvar. This procedure provides better flat embedding than
agarose. Up to three loops can be put in a single vial. Note: one does not
need super "EM" grade Formvar films, so this is not a hassle at all.
=A0=A0=A0=A0=A0=A0=A0When its time to embed, just cut off the stem of the l=
oop and
embed the loop with the sample on. The loop is heavy enough to sink and kee=
p
the root nice and flat. The copper is so thin you can cut through it with a=

razor in the block. Or if you like, you can excavate it from the block
and pull it out.
=A0=A0=A0=A0=A0=A0=A0Alternatively, if you don't want to mess with loops an=
d Formvar,
you can use agarose. Excise a 3mm tip segment and encase it within a small
droplet of 2% low gelling-temperature agarose (Type VII from Sigma). When
the agarose sets, the root-containing drop is transferred to 10% ethanol.
We get better results with Formvar but the agarose will work.
=A0=A0=A0=A0=A0=A0=A0The point of the loops or the agarose is to facilitate=
exchange of
solutions without loss of samples. It is a snap to suck out the old
solution and put in the new with your samples in agarose pellets or held on=
to
loops. With the agarose, you can keep all of the samples of a single
treatment in one vial (this saves time and solution) but with the formvar l=
oops,
it is hard to use more than three loops per vial.
=A0=A0=A0=A0=A0=A0=A0 I hope this helps. Good luck,=A0=A0=A0=A0=A0=A0=A0=A0=
=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0=A0Tobias
Baskin
p.s. I fix first. I am usually working with arabidopsis roots. We grow
the seedlings on agar plates and I pour the fixative right over the
seedlings on the plate. This way, they stay submerged and get nicely fixed.=
After
three rinses, ten min each, I then put the roots on the loops (or encase
them in agarose). Because the formvar is so thin, small molecules pass
through with no problem at all. We can even get antibodies through although=

these do get held up a bit.
-------
You can try eosin. We've used it to impart some color to early chicken
embryos that were later labelled by immuno-fluoresent markers. These
were
embedded im paraffin, and viewed in the LM, but it may cross-over.
Leona Cohen-Gould, M.S.
-----------
I have tried using fast green to lightly stain Arabidopsis roots (which
are
quite small) so I could still see them during embedding. I followed the
protocol
of Baskin et al., Planta 187: 405-413. They used methacrylate for
embedding
but it works as well in LR white. Just add 1 ul of 8% solution (w/v) fast
green dissolved in ethanol when your samples are at the 1st change to
100%
ethanol.
This should not interfere with immunolabeling. I also used acid fushin
but
have used this mostly in wax embedded material.
Elison Blancaflor
---------
Have you tried Eosin? Just add enough to color the fixative a nice
bright
pink, and process as usual. Another way to do it is to add make a 1.5%
solution of cobalt chloride in absolute ethanol, and using this as a
stock
solution, add 10 ml per 100 ml of absolute when processing. this colors
the
tiny stuff while dehydrating. Both of these methods are common in
routine
histology labs and do not interfere with immuno.
Wanda Shotsberger
--------
We've used both fast green and eosin Y in the 95% ethanol=20
step and been pleased with both. We used fast green to stain
isolated protoplasts.
Rosemary Walsh








From: Karen S Pawlowski :      kna101-at-utdallas.edu
Date: Wed, 8 Dec 1999 08:52:50 -0600 (CST)
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by ns0.utdallas.edu (Postfix) with ESMTP
id B788419FFF4; Wed, 8 Dec 1999 08:52:50 -0600 (CST)
Received: from localhost (kna101-at-localhost)
by apache.utdallas.edu (8.9.1/8.9.1) with SMTP id IAA10521;
Wed, 8 Dec 1999 08:52:50 -0600 (CST)
X-Authentication-Warning: apache.utdallas.edu: kna101 owned process doing -bs


I think Barbara has a good idea. Some e-mail systems can be set up to
automatically dump messages with a certain heading. To me, this would be
preferable to a total split.

Karen Pawlowski
Sr. Res. Assoc./UTSW Med. Ctr.
Dallas, TX

On Tue, 7 Dec 1999, Barbara Foster wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Mark,
}
} My early training came from the Royal Microscopical Society in the UK.
} They routinely meet with joint sessions, bio & mat sci. I have to tell you
} that each of these groups has a lot to learn from the other and would
} seriously discourage splitting of this group.
}
} An alternative: put an annotation in the header (ex: Bio: or MatS:.....).
} That way, those of us who are interested in learning on a broader scale
} can take advantage of the cross-pollination. Others who would like to stay
} more focused can just "clump and dump" the "uninteresting" messages.
}
} Best regards,
} Barbara Foster
} Consortium President
} Microscopy/Microscopy Education ...Educating microscopists for greater
} productivity.
}
} 125 Paridon Street Suite 102 Springfield, MA 01118
} PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
} Visit our web site {http://www.MME-Microscopy.com/education}
} ******************************************************
} MME is America's first national consortium providing
} customized on-site workshops in all areas of
} microscopy, sample preparation, and image analysis.
}
} At 08:04 AM 12/7/99 -0600, L. D. Marks wrote:
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} }
} } There are many advantages to having one general list for all
} } microscopies. However, there are also many disadvantages. My
} } primary interests are in the Physical Sciences, and I am not
} } interested in many of Biological topics that come up. As a
} } consequence the utility of this list for me has dropped a lot
} } as the volume of messages has increased.
} }
} } I think it is time to split this list into two, one for Biology
} } and one for the Physical Sciences. Of course, emails can still
} } be sent to both lists if the topic is appropriate enough.
} }
} } Comments?
} }
} } ++++++++++++++++++++++++++++++++++++++++++++++++
} } Laurence Marks
} } Department of Materials Science and Engineering
} } Northwestern University
} } fax: (847) 491-7820
} } mailto:l-marks-at-nwu.edu
} } http://www.numis.nwu.edu
} } ++++++++++++++++++++++++++++++++++++++++++++++++
} }
} }
} }
} }
}
}
}






From: Karen S Pawlowski :      kna101-at-utdallas.edu
Date: Wed, 8 Dec 1999 09:05:10 -0600 (CST)
Subject: Re: EM: background on slot grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Manuela,

Are the two types of grids you use made of the same metal? Some metals
can cause interferance-probably due to their charge. Are you
floating the grids on a drop of solution? If any solution gets to the
back side of the grid you could be doubling your background.

I haven't done this stuff for sometime-so my help in this area is limited.

Karen Pawlowski
SR. Res. Assoc./UTSW Med. Ctr.
Dallas, TX

On Wed, 8 Dec 1999, Palatsides, Manuela wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} We are having problems with terrible background when immunogold labelling
} sections on slot grids using either formvar or butvar as support films..
} Using the exact same procedure on 300 hexagonal grids with the same support
} film background is not a problem.
}
} The background in on the resin, all the cells and support film.
}
} Has anyone else experienced this problem using slot grids?
}
} Manuela Palatsides
} Electron Microscopy
} Peter MacCallum Cancer Institute
} Locked Bag#1
} A'Beckett Street
} Melbourne 3000
}
} Telephone: 03 96561244
} Fax: 03 96561411
} Email: m.palatsides-at-pmci.unimelb.edu.au
}
}
}
}






From: Douglas Keene :      DRK-at-SHCC.ORG
Date: Wed, 08 Dec 1999 07:12:06 -0800 (Pacific Standard Time)
Subject: Re: EM: background on slot grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Manuela,

You do not mention if your grids are made of the same
materials. Copper grids will oxidize during incubation in
immunocytochemical reagents, particularly those containing
azide (ie, commercial secondary gold conjugates), and also
in ammonium bicarbonate, among other buffers. Perhaps
there is a difference in the elements which compose your
grids?

----------------------
Douglas R. Keene
Associate Investigator
Shriners Hospital Research Unit
3101 S.W. Sam Jackson Park Road
Portland, Oregon 97201
phone: 503-221-3434
FAX: 503-412-6894 (9-5 PST)
e-mail: DRK-at-shcc.org







From: micro-at-ldeo.columbia.edu (Dee Breger)
Date: Wed, 8 Dec 1999 11:05:14 -0500
Subject: G (for general): dec boards needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,

My dear old Kevex 8000 is in need of vintage 1983 dec lsi 1123 circuit
boards: cpu board 8186, memory board 8044 and an un-numbered drive
interface board with a wide ribbon. Any help would be greatly appreciated!

And: my vote is for one list with subject line indicators.
(Great job, Nestor!!!)

Thanks,
Dee


Dee Breger
Mgr. SEM/EDX Facility
Lamont-Doherty Earth Observatory
61 Route 9W
Palisades, NY 10964 USA
T: 914/365-8640
F: 914/365-8155
E: micro-at-ldeo.columbia.edu

www.aspp.com/gallery/index_archive4.html
www.discovery.com/area/science/micro/micro1.html







From: Jo Dee :      jofish-at-burnham-inst.org
Date: Wed, 08 Dec 1999 08:18:44 -0800
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree wholeheartedly!!! The answers to many of the
questions on this list
benefit both the life and material science fields. The
questions and
answers about printers and cameras, etc. are most beneficial
to both fields.

On many mornings I download less than 20 messages, a number
easily handled
in just a few minutes. I tend to read all of my mail, even
though I may not
understand it all completely. I feel any new information is
helpful, and I
enjoy reading and learning about the "materials" side of
things as much as
the "biological" side.
I sincerely hope the listserver stays the same!
Jo Dee Fish

Gordon Couger wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} -From: L. D. Marks {ldm-at-apollo.numis.nwu.edu}
} }
} } There are many advantages to having one general list for all
} } microscopies. However, there are also many disadvantages. My
} } primary interests are in the Physical Sciences, and I am not
} } interested in many of Biological topics that come up. As a
} } consequence the utility of this list for me has dropped a lot
} } as the volume of messages has increased.
} }
} } I think it is time to split this list into two, one for Biology
} } and one for the Physical Sciences. Of course, emails can still
} } be sent to both lists if the topic is appropriate enough.
} }
}
} My interest are in LM and even less info is on this list about this.
} But I am very strongly for one list. It is a small list that generates
} less mail than I get on get rich quick spam every day. You don't
} have to read them all.
}
} The advantage to one list is the interdisplanary answeres to questions
} And this is the best list I have found for this.
}
} Please do FIX it is isn't broken.
}
} Gordon
} Gordon Couger
} 624 Cheyenne
} Stillwater, OK 74075
} 405 624 2855 GMT - 6:00

--
Jo Dee Fish
Electron Microscopy Assistant
The Burnham Institute
10901 N. Torrey Pines Rd.
La Jolla, CA 92037
858-646-3100 ext.3620







From: Keith Collins :      collins-at-alrc.doe.gov
Date: Wed, 8 Dec 1999 08:28:02 PST
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Well said Nestor. I agree if I do not want to read the message I can
delete it. Concerning headers well expect me to forgot.

Keith Collins
US DOE Albany Research Center.
Albany, Oregon





From: John F. Mansfield :      jfmjfm-at-engin.umich.edu
Date: Wed, 8 Dec 1999 12:51:39 -0500
Subject: General: List splitting and subject lines

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Yeah, big sigh here to!

I do echo the sentiment that the subject line should be as=20
descriptive as possible. I think we should all try to follow Joseph=20
Passero's example, he always identifies his subject clearly. Way to=20
go Joe!



At 6:59 PM -0600 12/7/99, L. D. Marks wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America



Please note new FAX number.

John Mansfield PhD CPhys MInstP
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (734) 936-3352 FAX (734) 763-2282 Cellular Phone: (734) 358-7555
(Leaving a phone message at 936-3352 is preferable to 358-7555)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html
Location: Lat. 42=B0 16' 48" Long. 83=B0 43' 48"





From: Lehman, Ann :      Ann.Lehman-at-trincoll.edu
Date: Wed, 8 Dec 1999 12:53:52 -0500
Subject: TEM of Aerogels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A colleague wishes to examine pore sizes and distribution in silica-based
aerogels by TEM. To thin the sample, we have tried glueing two aerogel
samples face to face, then tripod-polishing. We have tried increasing the
wedge angle, but still the samples break at the final polishing steps. We
have also tried to infiltrate the samples with resin that we cure prior to
polishing, to impart some mechanical strength to the pore regions. This too
has failed. The aerogels are susceptible to solubilization in acetone and
perhaps in polar resins.

We intend to try non-polar resins that are UV-polymerized and continue to
try the tripod polishing, and we may dope the resin with osmium (or other
mordant) to impart some contrast.

In the meantime, is anyone familiar with prepping this type of sample, or
references? Any suggestions?

Thanks
Ann Hein Lehman
Trinity College EM Facility





From: HILDEGARD CROWLEY :      hcrowley-at-du.edu
Date: Wed, 08 Dec 1999 10:58:13 -0700 (MST)
Subject: Which independent service contractor???

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Hi,

We are looking for an independent service company for our Hitachi 7000
TEM. We are considering Materials Analytical Services (Art McKenna),
Scientific Instrument Services (Alex Green), Vitaly Feingold Service
Company (Atlanta, GA), Pesto Inc., (Gynedd Valley, PA).

If anyone has experience with any of the above, we would so much
appreciate hearing from you.

Thank you,
Hildegard H. Crowley
University of Denver
Denver, CO
{hcrowley-at-du.edu}






From: Sara Miller :      saram-at-duke.edu
Date: Wed, 8 Dec 1999 14:09:11 -0500 (EST)
Subject: Re: TEM on viruses

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We do much TEM on viruses, both negative staining and thin sectioning.
We also get much material FedEx, so that it can be received very
quickly. If your customer is interested have him/her email, and I will
send details.

Sara Miller
(more address info below)

On Tue, 7 Dec 1999, Linda Rangell wrote:

} Date: Tue, 07 Dec 1999 16:25:46 -0800
} From: Linda Rangell {lkc-at-gene.com}
} To: MSA {Microscopy-at-sparc5.microscopy.com}
} Subject: TEM on viruses
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Does anyone know of a contract lab in the SF Bay Area that will do TEM
} of virus particles? I'm not sure if the person I'm asking for would
} like negative staining or thin sections.
}
} Thanks for any information,
} Linda Rangell
}
}
}

Sara E. Miller, Ph. D.
P. O. Box 3020
Duke University Medical Center
Durham, NC 27710
Ph: 919 684-3452
FAX: 919 684-8735






From: anderron-at-us.ibm.com
Date: Wed, 8 Dec 1999 14:28:42 -0500
Subject: Re: TEM of Aerogels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The advantage of tripod polishing is that you get a huge thin area. The
disadvantage of tripod polishing is that you get a huge thin area. This
translates to the fact that mechanically weak specimens don't have the
strength to hold together when thin over large areas.

Inasmuch as these areogel samples have a foam-like structure with brittle
cell walls and are more air than specimen, your idea of infiltrating the
samples with epoxy is correct. A low viscosity epoxy, like LR White works
well (in e.m. supplier catalogs, biological specimens) ((see what you learn
reading biological posts as well as phys sci posts!!)). If the epoxy
doesn't infiltrate far enough into the sample, then polish the first side
the best you can, paint epoxy onto the just-polished face, and make the
final first side polish. Soak in LR White to infiltrate into the polished
section and the second side polishing should see tha sample hold together.
Increasing the wedge angle is also correct--probably a full micrometer turn
or more. Try polishing in propylene glycol instead of water. Mount on a
small-hole grid.

If you still have problems, leave the specimen about 5 or 10 microns thick
and ion mill using a small diameter ion beam. The GATAN PIPS and similar
machines now have small beam ion guns that will create an ion milled dimple
less than 0.5 mm in size. This way the thicker material surrounding the
ion mill dimple provides strength to the thin region. We ***love*** those
small ion beams!


Ron Anderson, IBM, Hopewell Jct., New York, USA. anderron-at-us.ibm.com

Ann wrote:

A colleague wishes to examine pore sizes and distribution in silica-based
aerogels by TEM. To thin the sample, we have tried glueing two aerogel
samples face to face, then tripod-polishing. We have tried increasing the
wedge angle, but still the samples break at the final polishing steps. We
have also tried to infiltrate the samples with resin that we cure prior to
polishing, to impart some mechanical strength to the pore regions. This too
has failed. The aerogels are susceptible to solubilization in acetone and
perhaps in polar resins.

We intend to try non-polar resins that are UV-polymerized and continue to
try the tripod polishing, and we may dope the resin with osmium (or other
mordant) to impart some contrast.

In the meantime, is anyone familiar with prepping this type of sample, or
references? Any suggestions?







From: 7news4u1-at-pastunsn.net
Date: Sun, 09 Jan 2000 00:30:36 -0800
Subject: Get Windows 98 Second Edition Update ?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


{HTML}
{BODY}

{FONT face=3D"MS Sans Serif"}
{FONT size=3D2}
{FONT color=3D"#0000FF"} {B} Here's the December 8 KTS Wholesale Software M=
embers Update {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
**************************************************************************=
**** {BR}
{BR}
If you've been wanting {/B} {/FONT}
{FONT color=3D"#FF0000"} {B} software for Christmas {/B} {/FONT}
{FONT color=3D"#000000"} {B} to give as gifts or for {BR}
yourself, look at the special offers we have. All your favorites and {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {U} ALL deliverable before Christmas {/U} {/B} {/F=
ONT}
{FONT color=3D"#000000"} {B} ! {BR}
{BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} MS Office 2000 Professional Full vers=
ion CDs only $ 189.95 ** {BR}
{LI} MS Windows 98 Second Edition Full version $ 94.95 ** {BR}
{LI} MS Windows 98 Second Edition Update $ 39.95 ** {BR}
{LI} MS Outlook 2000 Full version $ 67.95 ** {BR}
{/LI} {/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Symantec Act 2000 Full Version, No re=
bate hassles ! $ 99.95 {BR}
{LI} Norton Utilities 2000 $ 22.95 * {BR}
{LI} Norton Antivirus 2000 $ 24.95 * Retail Box Set {BR}
{/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} McAfee Office Suite w First Aid 2000,=
! a $ 69.99 value only $ 22.95 {BR}
{LI} McAfee VirusScan v 4.02, ! a $ 39.99 value only $ 9.95 {BR}
{/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Lotus Smartsuite Millenium Ed. v 9.0 =
$ 69.95 * Compare to $379.00 {BR}
{LI} Corel Wordperfect Office 2000 Full version $ 69.95 No rebate hassles =
! {BR}
{/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Microsoft NT Server 4.0 w 5 CAL's $ 4=
99.95 {BR}
{LI} Microsoft NT Workstation v. 4.0 $ 94.95 {BR}
{LI} Microsoft Backoffice Server v 4.5 Small Bus. Ed w 5 CALs $ 990.95 {BR}
{/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Adobe Pagemaker 6.52 Plus Retail Box =
set $ 315.95 {BR}
{LI} Adobe Photoshop 5.5 Full Version $ 389.95 * NEW !! {BR}
{LI} Adobe InDesign 1.0 Full Version $ 489.95 * NEW !! {BR}
{/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Macromedia Dreamweaver v 3.0 NEW !!! =
$ 207.95 {BR}
{LI} Macromedia Drumbeat 2000 NEW !!! $ 207.95 {BR}
{/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{BR}
{BR}
Other products are available as well. {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {BR}
For more information or to order, please visit our SECURE web site. {BR}
{BR}
{/FONT}
{FONT color=3D"#FF0000"} Order now ! Guaranteed delivery by Christmas ! {BR=
}
{/FONT}
{FONT color=3D"#000000"} {BR}
{a href=3D"http://www.cdsales.net/cgi-bin/index.cgi?1829032"} {font size=3D=
2} {center} {/FONT}
{FONT color=3D"#0000FF"} {B} Click Here to Order or {br} see other Special O=
ffers of Great Software {BR}
www.cdsales.net {/B} {/FONT}
{FONT color=3D"#000000"} {/center} {/a} {BR}
{BR}
Or if the above link isn't active, cut and paste this into your browser: {B=
R}
{BR}
{/FONT}
{FONT color=3D"#0000FF"} http://www.cdsales.net/cgi-bin/index.cgi?1829032 {=
/FONT}
{FONT color=3D"#000000"} {BR}
{BR}
Thank you and have a great day. {BR}
{BR}
{center} ******************************************************************=
*************************** {br} These Announcements are periodic notices o=
f new features, special events and free offers available to members and su=
bscribers. {FONT face=3D"Arial"} {p} To Be Removed , Click Here {a href=3D=
mailto:removes-at-jh1.net?subject=3Dremove} {BR} Remove Me {/a} {br} ***********=
**************************************************************************=
******** {/center} {BR}
{BR}
{/FONT} {/FONT} {/BODY} {/HTML}







From: 4newsletter2-at-pastunsn.net
Date: Sun, 09 Jan 2000 00:39:43 -0800
Subject: Get Windows 98 Second Edition Update ?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


{HTML}
{BODY}

{FONT face=3D"MS Sans Serif"}
{FONT size=3D2}
{FONT color=3D"#0000FF"} {B} Here's the December 8 KTS Wholesale Software M=
embers Update {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
**************************************************************************=
**** {BR}
{BR}
If you've been wanting {/B} {/FONT}
{FONT color=3D"#FF0000"} {B} software for Christmas {/B} {/FONT}
{FONT color=3D"#000000"} {B} to give as gifts or for {BR}
yourself, look at the special offers we have. All your favorites and {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {U} ALL deliverable before Christmas {/U} {/B} {/F=
ONT}
{FONT color=3D"#000000"} {B} ! {BR}
{BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} MS Office 2000 Professional Full vers=
ion CDs only $ 189.95 ** {BR}
{LI} MS Windows 98 Second Edition Full version $ 94.95 ** {BR}
{LI} MS Windows 98 Second Edition Update $ 39.95 ** {BR}
{LI} MS Outlook 2000 Full version $ 67.95 ** {BR}
{/LI} {/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Symantec Act 2000 Full Version, No re=
bate hassles ! $ 99.95 {BR}
{LI} Norton Utilities 2000 $ 22.95 * {BR}
{LI} Norton Antivirus 2000 $ 24.95 * Retail Box Set {BR}
{/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} McAfee Office Suite w First Aid 2000,=
! a $ 69.99 value only $ 22.95 {BR}
{LI} McAfee VirusScan v 4.02, ! a $ 39.99 value only $ 9.95 {BR}
{/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Lotus Smartsuite Millenium Ed. v 9.0 =
$ 69.95 * Compare to $379.00 {BR}
{LI} Corel Wordperfect Office 2000 Full version $ 69.95 No rebate hassles =
! {BR}
{/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Microsoft NT Server 4.0 w 5 CAL's $ 4=
99.95 {BR}
{LI} Microsoft NT Workstation v. 4.0 $ 94.95 {BR}
{LI} Microsoft Backoffice Server v 4.5 Small Bus. Ed w 5 CALs $ 990.95 {BR}
{/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Adobe Pagemaker 6.52 Plus Retail Box =
set $ 315.95 {BR}
{LI} Adobe Photoshop 5.5 Full Version $ 389.95 * NEW !! {BR}
{LI} Adobe InDesign 1.0 Full Version $ 489.95 * NEW !! {BR}
{/LI} {/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {B} {UL} {LI} Macromedia Dreamweaver v 3.0 NEW !!! =
$ 207.95 {BR}
{LI} Macromedia Drumbeat 2000 NEW !!! $ 207.95 {BR}
{/LI} {/LI} {/UL} {/B} {/FONT}
{FONT color=3D"#000000"} {B} {BR}
{BR}
{BR}
Other products are available as well. {BR}
{/B} {/FONT}
{FONT color=3D"#000000"} {BR}
For more information or to order, please visit our SECURE web site. {BR}
{BR}
{/FONT}
{FONT color=3D"#FF0000"} Order now ! Guaranteed delivery by Christmas ! {BR=
}
{/FONT}
{FONT color=3D"#000000"} {BR}
{a href=3D"http://www.cdsales.net/cgi-bin/idx1.cgi?7566430"} {font size=3D=
2} {center} {/FONT}
{FONT color=3D"#0000FF"} {B} Click Here to Order or {br} see other Special O=
ffers of Great Software {BR}
www.cdsales.net {/B} {/FONT}
{FONT color=3D"#000000"} {/center} {/a} {BR}
{BR}
Or if the above link isn't active, cut and paste this into your browser: {B=
R}
{BR}
{/FONT}
{FONT color=3D"#0000FF"} http://www.cdsales.net/cgi-bin/ {/FONT}
{FONT color=3D"#000000"} idx1.cgi?7566430 {BR}
{BR}
Thank you and have a great day. {BR}
{BR}
{center} ******************************************************************=
*************************** {br} These Announcements are periodic notices o=
f new features, special events and free offers available to members and su=
bscribers. {FONT face=3D"Arial"} {p} To Be Removed , Click Here {a href=3D=
mailto:removes-at-jh1.net?subject=3Dremove} {BR} Remove Me {/a} {br} ***********=
**************************************************************************=
******** {/center} {BR}
{BR}
{/FONT} {/FONT} {/BODY} {/HTML}







From: Shotsberger-Gray, Wanda :      WandaShotsberger-Gray-at-hmhs.com
Date: Tuesday, December 07, 1999 11:10AM
Subject: Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree with Barbara, one list for all with coding if the consensus goes
that way.
Wanda
----------
} From: Barbara Foster
To: L. D. Marks; Microscopy List
-----------------------------------------------------------------------.


Mark,

My early training came from the Royal Microscopical Society in the UK.
They routinely meet with joint sessions, bio & mat sci. I have to tell you
that each of these groups has a lot to learn from the other and would
seriously discourage splitting of this group.

An alternative: put an annotation in the header (ex: Bio: or MatS:.....).
That way, those of us who are interested in learning on a broader scale
can take advantage of the cross-pollination. Others who would like to stay
more focused can just "clump and dump" the "uninteresting" messages.

Best regards,
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************
MME is America's first national consortium providing
customized on-site workshops in all areas of
microscopy, sample preparation, and image analysis.

At 08:04 AM 12/7/99 -0600, L. D. Marks wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America






From: Soumitra Ghoshroy :      ghoshroy-at-nmsu.edu
Date: Wed, 08 Dec 1999 14:23:47 -0700
Subject: Thank you

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,

Thank you all very much for responding to my inquiry about good books on
microscopy. Here are the responses I received.

1. Microscopy and Photomicrography - A Working Manual
Robert F. Smith, CRC Press, Inc., Boca Raton, Florida, 1990.
ISBN - 0-8493-8803-1

2. Light Microscopy in Biology - A Pratical Approach
Alan J. Lacey, IRL Press, New York, 1991.
ISBN - 0-19-963036-4 (hardbound)
ISBN - 0-19-963037-2 (softbound)

The second book contains a special Flourescence Microscopy by P.S. Ploem.


O=B4Brian and McCully (1981): The study of plant structure. Principles and
selected methods. Merlbourne Australia.
old but good


Light and Electron Microscopy by Elizabeth M. Slayter & Henry S. Slayter.
It does mention fluorenscence microscopy but not in great details, but very
good of light microscopy in general.


"Video Microscopy" by Sluder & Wolf or "Green Fluorescent
Proteins" by Sullivan and Kay. Both available from Academic Press
www.apnet.com or "Fluorescence Microscopy" (Microscopy Handbooks, 40); H. J.
Tanke, Brian Herman www.amazon.com


"Optimizing Light Microscopy for Biological and Clinical Laboratories" was
written just for this purpose. Jim Pawley has been using it at both U. WI
and UBC; Dave Knecht uses it at UConn. Has learning objectives at the
beginning of each chapter as well as short quizzes at the end. Covers both
basic principles and has a detailed section on fluorescence... even a bit
on confocal.
Classroom discounts are also available. Please see
MME-Microscopy.com/education for further details.


Soumitra Ghoshroy Ph.D.
Electron Microscope Laboratory
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Phone: 505-646-3600/1531
Fax: 505-646-5665
E-mail:ghoshroy-at-nmsu.edu





From: Caroline Schooley :      schooley-at-mcn.org
Date: Wed, 8 Dec 1999 13:38:47 -0800 (PST)
Subject: Project MICRO

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Project MICRO (Microscopy In Curriculum - Research Outreach) is MSA's
middle school outreach program. Many of you have heard of it, but you may
not have visited its website (URL below). Now would be a good time,
because Nestor, in his abundant spare time, has just posted a major
revision. You'll find up-to-date reviews of lots of books, videos &
CD-ROMs about microscopy. Most, but not all, are for children; you may get
some good holiday gift ideas. There is also a much-expanded website
hotlink list. If you are aware of anything that has been missed, please
let me know.

I also need potential sources of funding for microscopes for schools. I'm
looking for a foundation or program that might be interested in an
equipment-only proposal at the $40-50K level. Suggestions?

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html







From: Mel Dickson :      m.dickson-at-unsw.edu.au
Date: Thu, 09 Dec 1999 10:06:33 +1100
Subject: Kevex LSI-11 spares

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Like Dee Breger, we have a Kevex (Delta-Plus) EDS which has a fault in the
LSI-11 system.

Any replacement boards which might be donated or sold cheap, (cpu board
8186, memory board 8044 and an un-numbered drive) would be most welcome.





From: jcoleman-at-rdg.boehringer-ingelheim.com
Date: Wed, 8 Dec 1999 18:32:03 -0600
Subject: One List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


For all the reasons cite; especially the excellent stewardship of Nestor;
one list is my choice.







From: Glenn Poirier :      glennp-at-eps.mcgill.ca
Date: Wed, 8 Dec 1999 18:32:50 -0600
Subject: Re: Time for more than one list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Don't mess with success!

With a unified list like this I've been exposed to great ideas from people
that I would not have otherwise talked to.

The volume of mail is not that large that good subject lines and the delete
button can't handle it.

Glenn


=============================================================
Glenn Poirier Tel:(514) 398-6774
MicroAnalytical Laboratory Fax:(514) 398-4680
Earth and Planetary Sciences
McGill University glennp-at-eps.mcgill.ca
3450 University St. castaing.eps.mcgill.ca
Montreal, Qc
H3A 2A7 -- Millennium Hand and Shrimp --
==============================================================







From: uri :      uri-at-watson.ibm.com
Date: Wed, 8 Dec 1999 18:39:01 -0600
Subject: Re: new list ideas

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Laurie Wallin says:
} I agree with Barbara Foster's comment. I am working in a branch of biology
} and though I do not understand the whole of the physical microscopy
} workings, I enjoy hearing about interesting topics no matter what field. I
} like the idea to attach a heading to the front of subject lines that would
} designate messages as Bio, versus Mat. Sci.

Since somebody apparently is counting the opinions (votes?), here's
mine.

Splitting the list is a BAD idea, and does not make much sense.
Plus, upon what criteria would you have it split? I personally
prefer to split by LM vs. EM... See the point...?
--
Regards,
Uri uri-at-watson.ibm.com
-=-=-=-=-=-=-
{Disclaimer}







From: BrosnanWatters, Gayle :      GBrosnanWatters-at-vanguard.edu
Date: Wed, 8 Dec 1999 18:39:40 -0600
Subject: splitting the list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I tend to agree with Barbar and Laurie - and even more so for myself! I am
a biological psychologist and the only connection I even have with the group
is that I cut mouse brains on an ultramicrotome and then look at them with
very simple light microscopy. However, I learn something from some
postings, and what I don't want to read I delete - I would be afraid I'd
miss something I DO need if you had two lists.
Thanks!!







From: Randy & Jenna & Orin Brown :      randyjen-at-northnet.org
Date: Wed, 8 Dec 1999 18:42:48 -0600
Subject: split up the list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am only a student member of the MSA, and so I find the discussions
on all subjects potentially valuable!  So far, my interest is
primarily SEM, but I am not sure where my career may take me... just my
1.2 cents worth (students don't have a whole 2 cents to spare) Jenna
Brown student, SUNY Potsdam (hi Dr. Rhoads!)







From: Peter Bond :      P.Bond-at-plymouth.ac.uk
Date: Wed, 8 Dec 1999 18:47:44 -0600
Subject: Cryo-SEM and failed delivery

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry to have to post this to the list ( although it's a good advert
anyway) but my off-list message to Tian did not get there.

{"Tian_Huang-at-gillette.com"-at-sparc5.microscopy.com}

So

Tian

If you're in th UK we can do cryo-SEM in our lab.

Let me know what you are looking at, and we can discuss your work.

Regards

Pete

Peter Bond
Plymouth Electron Microscopy Unit
University of Plymouth
Drake Circus
Plymouth
Devon
UK PL4 8AA
Tel/Fax: 01752 233092
email: pbond-at-plymouth.ac.uk







From: Palatsides, Manuela :      m.palatsides-at-pmci.unimelb.edu.au
Date: Thu, 9 Dec 1999 12:45:00 +1100
Subject: EM: Background on slot grids

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry I didn't mention that we use Nickel grids for all our immunolabelling,
so that's not the problem.

Manuela Palatsides
Electron Microscopy
Peter MacCallum Cancer Institute
Locked Bag#1
A'Beckett Street
Melbourne 3000

Telephone: 03 96561244
Fax: 03 96561411
Email: m.palatsides-at-pmci.unimelb.edu.au







From: p00bare :      p00bare-at-pdq.net
Date: Wed, 08 Dec 1999 22:28:52 -0600
Subject: Re: TEM of Aerogels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Is the aerogel closed-cell or open-cell? Closed cell (like styrofoam)
is impermeable and won't absorb much resin. Dave Pevear, Houston

"Lehman, Ann" wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} A colleague wishes to examine pore sizes and distribution in silica-based
} aerogels by TEM. To thin the sample, we have tried glueing two aerogel
} samples face to face, then tripod-polishing. We have tried increasing the
} wedge angle, but still the samples break at the final polishing steps. We
} have also tried to infiltrate the samples with resin that we cure prior to
} polishing, to impart some mechanical strength to the pore regions. This too
} has failed. The aerogels are susceptible to solubilization in acetone and
} perhaps in polar resins.
}
} We intend to try non-polar resins that are UV-polymerized and continue to
} try the tripod polishing, and we may dope the resin with osmium (or other
} mordant) to impart some contrast.
}
} In the meantime, is anyone familiar with prepping this type of sample, or
} references? Any suggestions?
}
} Thanks
} Ann Hein Lehman
} Trinity College EM Facility






From: jim :      jim-at-proscitech.com.au
Date: Thu, 9 Dec 1999 11:41:25 +1000
Subject: RE: TEM of Aerogels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Perhaps putting a keyword in the subject heading would allow individuals to
separate listserv messages into separate files or delete them automatically
through functions in e-mail programs. I would like to preserve a commensal
relationship.
-Karl Garsha
----- Original Message -----
} From: {"r.cross-at-ru.ac.za"-at-sparc5.microscopy.com}
To: {microscopy-at-sparc5.microscopy.com}
Sent: Tuesday, December 07, 1999 11:27 PM


Surface replicas would be more certain to work; for these you may require a
good polish so that the pore distribution is more accurate. A polished (at
least a reasonably flat) surface would be easier to produce, because it can be
thick and strong. Plastic replicas, because of the solvent requirement would
not be possible, but carbon/platinum replicas should work. You probably would
dissolve the specimen for cleaning purposes.

If FESEM gives enough resolution (and is available), this could save a heap of
trouble.

Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Thursday, December 09, 1999 3:54 AM, Lehman, Ann
[SMTP:Ann.Lehman-at-trincoll.edu] wrote:
}
}
} A colleague wishes to examine pore sizes and distribution in silica-based
} aerogels by TEM. To thin the sample, we have tried glueing two aerogel
} samples face to face, then tripod-polishing. We have tried increasing the
} wedge angle, but still the samples break at the final polishing steps. We
} have also tried to infiltrate the samples with resin that we cure prior to
} polishing, to impart some mechanical strength to the pore regions. This too
} has failed. The aerogels are susceptible to solubilization in acetone and
} perhaps in polar resins.
}
} We intend to try non-polar resins that are UV-polymerized and continue to
} try the tripod polishing, and we may dope the resin with osmium (or other
} mordant) to impart some contrast.
}
} In the meantime, is anyone familiar with prepping this type of sample, or
} references? Any suggestions?
}
} Thanks
} Ann Hein Lehman
} Trinity College EM Facility






From: Gordon Couger :      gcouger-at-rfdata.net
Date: Thu, 9 Dec 1999 00:03:07 -0600
Subject: Re: Project MICRO

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



-----Original Message-----
} From: Caroline Schooley {schooley-at-mcn.org}
}
}
} Project MICRO (Microscopy In Curriculum - Research Outreach) is MSA's
} middle school outreach program. Many of you have heard of it, but you may
} not have visited its website (URL below). Now would be a good time,
} because Nestor, in his abundant spare time, has just posted a major
} revision. You'll find up-to-date reviews of lots of books, videos &
} CD-ROMs about microscopy. Most, but not all, are for children; you may get
} some good holiday gift ideas. There is also a much-expanded website
} hotlink list. If you are aware of anything that has been missed, please
} let me know.
}
} I also need potential sources of funding for microscopes for schools. I'm
} looking for a foundation or program that might be interested in an
} equipment-only proposal at the $40-50K level. Suggestions?
}
Caroline ,

If it is for equipment I would put the arm on the manufactures. They can
claim
retail as a tax deduction and 30% of retail will cover their cost of
production so
it is not costing them money. In fact they may make a little. The markup is
pretty
high in slow turn over items like microscopes.

For those not familiar with the US tax structure the tax rate is from 25 to
34% of
profit. So if the tax rate is 30% a 100 dollar tax deduction will save 30
dollars in
taxes. If it cost 35 dollars to make an item that retails for 100 and you
donate it
to a school or other non profit entity you can deduct it from your profit
reducing
you income tax by $30. The actual cost to the company is 5$

Gordon

Gordon Couger gcouger-at-couger.com

Stillwater, OK www.couger.com/gcouger
405 624-2855 GMT -6:00 www.couger.com/gcouger







From: Radostin Danev :      rado-at-nips.ac.jp
Date: Thu, 9 Dec 1999 15:24:21 +0900
Subject: PHY. Re: Time for more than one list?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I vote for one list. The reasons have been given and repeated many times
already.

This scintillation became a chain-reaction. :-) (the physics part)

Best wishes,

Rado

---------------------------------------------------------------------
Radostin Danev
Laboratory of Ultrastructure Research
National Institute for Physiological Sciences
Myodaiji-cho, Okazaki 444-8585, JAPAN
e-mail: rado-at-nips.ac.jp
---------------------------------------------------------------------






From: Krzysztof Jan Hubner :      hubner-at-IOd.krakow.pl
Date: Signature
Subject: Conference Eutectica V

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



INTERNATIONAL CONFERENCE
EUTECTICA V
12 - 14 June 2000 , Dniepropetrovsk, Ukraine

ORGANIZING COMMITTEE

National Metallurgical Academy,
ave. Gagarin, 4, Dniepropetrovsk, 5, 49600, Ukraine
Tel: +380 562 410 602; Tel., Fax :+380 562 676 977;
E-mail:mazur-at-dmeti.dp.u
EUTECTICA-V International Scientific Conference. 12-14 June 2000

FIRST CIRCULAR

ORGANIZERS
ASM International
National Metallurgical Academy of Ukraine
Division of Materials Science and Metallurgy, Academy of Engineering
Sciences of Ukraine

ORGANIZING COMMITTEE
Chairman: Prof. Yurii N. Taran, Member, National Academy of Sciences of
Ukraine
Co-Chairman: Dr. Hans Portisch, President, ASM International
Vice-Chairmen: Prof. Vladyslav I. Mazur.
Phone: +380 562 41 06 02.
Tel./fax: +380 562 67 69 77.
Prof. Vladimir I. Shapovalov.
Phone: +1 505 275 1625.
Secretary: Dr. Svetlana V. Kapustnikova.
Phone: +380 562 41 06 02.

ADDRESS
National Metallurgical Academy of Ukraine
Prospekt Gagarina, 4
Dniepropetrovsk 320635, Ukraine
Tel./fax: +380 562 67 69 77. E-mail: mazur-at-dmeti.dp.ua

SCIENTIFIC TOPICS
1. Atomic structure of eutectic melts
2. Mechanism and atomic kinetics of eutectic solidification
3. Influence of cluster structure in melt on kinetics of eutectic
solidification
4. Directional solidification of eutectics
5. Three dimensional models of eutectic grains
6. Inoculation and modification of eutectic alloys
7. Thermal stability of eutectic alloys
8. Eutectic alloys properties
9. New eutectic type alloys

CONTRIBUTOR DATA
Name (last, first, middle):
Academic degree: Affiliation:
Paper title:
Paper language (please underline): Ukrainian English Russian
Mailing address:
Contact phone:
Fax, e-mail

GUIDELINES
Your abstract should be submitted in hard copy on A4 paper and as a
separate file on the floppy disk. Please follow the Instructions.

INSTRUCTIONS TO AUTHORS
Margins: Top, 2.0 cm. Left, 2.5 cm. Right, 2.5 cm. Bottom, 3.0 cm.
Font: Arial or Arial Cyr 14 pt. Spacing: 1 1/2. Justification: left and
right. Print resolution: 300 dpi or more. Please use a laser, a bubble or a
24-pin dot matrix printer.

PAPER TITLE
X. Namea; Y. Nameb
a) Affiliation. City. -Country.
b) Affiliation. City. -Country.
(Blank line)
Text up to three full pages in A4 format.


Please use Word for Windows 6.0 or 7.0. For each photo provide a separate
file in TIFF or CDR(5.0) and for each simple drawing, a PCX or WMF file.
Your floppy disk is returned on request.

CONFERENCE FEE
The conference fee covers the costs of publishing the conference
proceedings, servicing the attendees, and a final dinner. All abstracts are
included in the abstracts volume after receipt by the Organizing Committee
of the conference fee. The fee is equivalent to USD 50, if received not
later than 1 May 2000 or USD 55 if paid later.

FEE REMITTANCE ADDRESS
Mail order: Account ?000071293 LATVIAN TRADE BANK TRIJADIBAS St.,4, RIGA,
LV-1048, LATVIA ABN AMRO BANK N.Y., NEW YORK, USA
Cor.Acc. ?574072285541, SWIFT: ABNA US 33
Please mark EUTECTICA-V on your order form.

ADDITIONAL INFORMATION
1. The Organizing Committee will appreciate your forwarding this form to
your colleagues who might be interested in attending the Conference.
2. Space for sponsors' advertisements will be provided in the Conference
Proceedings. The Organizing Committee invites all interested parties to
send in their advertising materials.







From: pe13-at-cam.ac.uk
Date: Thu, 09 Dec 1999 10:06:20 +0000
Subject: Please, only one list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by ursa.cus.cam.ac.uk with esmtp (Exim 3.12 #1)
id 11w0Rg-0000BS-00; Thu, 09 Dec 1999 10:05:09 +0000


The important and interesting advances in science occur at the interfaces
of different disciplines. Please keep one list, for as a biologist I am
interested to read what non-biologists do in microscopy and analysis. It
may mean going through 100 messages to find 1 gem of a new idea, but it's
worth it.

Patrick Echlin
Cambridge UK







From: Garber, Charles A. :      cgarber-at-2spi.com
Date: Thu, 09 Dec 99 05:13:32 -0500
Subject: TEM of aerogels

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


-- [ From: Garber, Charles A. * EMC.Ver #3.1 ] --

Ann Hein Lehman wrote:
==============================================================
A colleague wishes to examine pore sizes and distribution in silica-based
aerogels by TEM. To thin the sample, we have tried glueing two aerogel
samples face to face, then tripod-polishing. We have tried increasing the
wedge angle, but still the samples break at the final polishing steps. We
have also tried to infiltrate the samples with resin that we cure prior to
polishing, to impart some mechanical strength to the pore regions. This too
has failed. The aerogels are susceptible to solubilization in acetone and
perhaps in polar resins.

We intend to try non-polar resins that are UV-polymerized and continue to
try the tripod polishing, and we may dope the resin with osmium (or other
mordant) to impart some contrast.

In the meantime, is anyone familiar with prepping this type of sample, or
references? Any suggestions?
================================================================
These kinds of samples can be prepared much the same way as a refinery
catalyst, or a zeolite system, that is, using

a) standard TEM embedding resins, and
b) diamond knife ultramicrotomy

There are several "must do's" in the technique, realizing also that they can
be accomplished differently by different users:

• Relatively "hard" block is needed because this is a "hard" sample.
We usually would use our own SPI-Pon 812 resin, but some of the other "Epon
substitutes" would work, we assume, just as well. I am not aware that we
have had such samples dissolving in this resin. There should be enough
electron contrast, assuming the sections are thin enough.

• You want to have excellent infiltration, to the degree that (we
believe) is generally not possible (at least with certainty) without the use
of vacuum impregnation approaches.

• Diamond knife ultramicrotomy, using a "materials science" diamond
knife. Naturally we prefer our own SPI diamond knives, but in this case,
whatever knife you do use, we believe a 45° angle edge is better than a 55°
edge (for this kind of work).

• This has to be done on a good ultramicrotome, and in the hands of
an experienced person. One also has to be aware of artifacts, as can occur
in any technique, but the principal artifacts here are knife marks and
possibly particles pulled out with the knife. But these are easily
recognizable as artifacts where as artifacts caused during some of the other
potential techniques tend to be more isotropic in nature and therefore more
difficult to identify as artifacts.

Chuck

============================================

Charles A. Garber, Ph. D. Ph: 1-610-436-5400
President 1-800-2424-SPI
SPI SUPPLIES FAX: 1-610-436-5755
PO BOX 656 e-mail:cgarber-at-2spi.com
West Chester, PA 19381-0656 USA
Cust.Service: spi2spi-at-2spi.com

Look for us!
########################
WWW: http://www.spi.cc
########################
============================================





From: Witold Zielinski :      wiziel-at-inmat.pw.edu.pl
Date: Thu, 9 Dec 1999 12:52:45 CET
Subject: Re: One List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since it seems that it is a kind of voting - I vote to keep one list
(despite that I am interested with metals and crystals).
Cheers,
Witold Z.

'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''

Witold Zielinski, Ph.D.
Warsaw University of Technology
Department of Materials Science and Engineering
Narbutta 85, 02 524 Warszawa
POLAND

phone: (48 22) 660 84 46
fax: (48 22) 48 48 75





From: pogany-at-power.szfki.kfki.hu (Pogany Lajos)
Date: Thu, 9 Dec 1999 07:56:09 -0600
Subject: Re: One List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I found that there are a lot of usfeful information in all letters and
there is not reasonable to split the newsgroup. A lot of informations are
interesting to me in the biology despite that I am interested with metals
and magnetism, so I would like to remain at only one group.
Cheers,
L. Pogany









From: pjrandsjr-at-worldnet.att.net
Date: Thu, 9 Dec 1999 08:05:45 -0600
Subject: GEN: Electron microscope as a turning point in history?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues... This is a broad enough question that it can have many
answers. Would any of you like to help me answer this for this student?

I would think answers to both the list and to this student would be useful
since he is not a member of the Server.

Nestor
Your Friendly Neighborhood SysOp
-----------------------------------------------------------------------

Email: pjrandsjr-at-worldnet.att.net
Name: Philip Randolph
School: Gonzaga Preparatory School


Question: We are looking into the development of the electron microscope as
a turning point in history. What specific information can you provide that
will give us a sense of the effect of the advent of the electron microscope
on the cultural, social and economic environment of society? Thank you for
your time in considering this question.








From: Tigran Dolukhanyan :      Tigran_Dolukhanyan-at-uml.edu
Date: Thu, 09 Dec 1999 10:37:05 -0500
Subject: One list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am strongly for ONE LIST ONLY.






From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 09 Dec 1999 10:32:23 -0500
Subject: Re: film dryers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Leona Cohen-Gould wrote:

} We are in the market for a small film drying cabinet for our EM film. We
} don't need, or have space for one of thoses locker-sized jobs they sell in
} photo suplly catalogs. Any ideas on where to look? Or better still, is
} anyone looking to get rid of one?
}

Dear Lee,
In the darkroom where we don't have a wall-mounted film dryer, we

use a lab oven at 45 C. The wall-mounted dryers are about 36" high, 24"
wide, and 18" deep. They fit nicely over the sink. My personal preference
is to air-dry the film overnight; there's less spotting and streaking that
way.
Yours,
Bill Tivol






From: Soumitra Ghoshroy :      ghoshroy-at-nmsu.edu
Date: Thu, 09 Dec 1999 08:33:14 -0700
Subject: Job opening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


ELECTRON MICROSCOPY SPECIALIST
(Materials Science)
New Mexico State University


The Electron Microscopy Laboratory at New Mexico State University is
seeking an Electron Microscopy Specialist in Materials Science area. The
laboratory provides transmission and scanning electron microscopy and some
light microscopy services for the university research community and a few
external organizations, in biological, physical and materials science fields.

Qualifications: Bachelors degree in a physical science or materials
engineering area (in hand by hire date). Masters degree desirable.
Candidates must have at least three years of related experience in scanning
electron microscopy and energy-dispersive X-ray microanalysis. TEM
experience would be a plus. A working knowledge of Windows based PC
computers is essential. Experience with digital image capture and
analysis, fluorescence microscopy is desirable. The candidate must be
familiar with materials sample preparation techniques, vacuum evaporation,
sputter coating, mechanical and electronic equipment, and vacuum systems.
The successful candidate must be able to work well with researchers, staff,
and students and be able to train graduate and undergraduate students for
independent work with relevant techniques and equipment. The candidate
should be able to work independently, plan and organize multiple projects
and willing to learn new techniques.

Duties and Responsibilities: Operation and routine maintenance of
transmission and scanning electron microscopes and associated equipment;
supervision of facility users; record keeping, including billings, budgets,
toubleshooting and maintenance of instrument and research logs, specimen
preparation, data generation and some analysis.

Salary: DOQ Website: www.nmsu.edu/~personal/postings/professional/

Screening of applicants will begin January 4, 2000 and continue until a
candidate is chosen.

Applications should include a resume, letter of application and three
letters of recommendation. Position contingent upon funding. NMSU is an
EEO/AA Employer.

Apply to:
Dr. Reed Dasenbrock
Associate Dean/Director
Arts & Sciences Research Center
New Mexico State University
MSC RC, Box 30001
Las Cruces, NM 88003
cgower-at-nmsu.edu


Soumitra Ghoshroy Ph.D.
Electron Microscope Laboratory
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Phone: 505-646-3600/1531
Fax: 505-646-5665
E-mail:ghoshroy-at-nmsu.edu





From: William Tivol :      tivol-at-wadsworth.org
Date: Thu, 9 Dec 1999 10:41:33 -0500 (EST)
Subject: Gen: Chemical resistance of gloves

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The LabSource catalog has a chemical resistance chart for latex
and nitrile gloves and about a dozen pages of latex, nitrile and vinyl
gloves for sale. I have no affiliation with this company.
Bill Tivol





From: Michael Bode :      mb-at-soft-imaging.com
Date: Thu, 9 Dec 1999 08:37:35 -0700
Subject: RE: One List -- stop this thread?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Is anybody keeping track of these messages?

If yes, could that person identify her/himself so that these messages
can be directed directly there?

If not, I think it is pretty evident that the majority is for keeping
one list and I think we can safely stop this thread and use the
bandwidth for other information.

(Please don't start another thread now about whether we can stop this or
not!!!!)

Just my humble opinion....

Michael

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================


} ----------
} From: Witold Zielinski[SMTP:WIZIEL-at-INMAT.PW.EDU.PL]
} Sent: Thursday, December 09, 1999 5:52:45 AM
} To: Microscopy-at-sparc5.microscopy.com;
"jcoleman-at-rdg.boehringer-ingelheim.com"-at-sparc5.microscopy.com
} Subject: Re: One List
} Auto forwarded by a Rule
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Since it seems that it is a kind of voting - I vote to keep one list
(despite that I am interested with metals and crystals).
Cheers,
Witold Z.

'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''

Witold Zielinski, Ph.D.
Warsaw University of Technology
Department of Materials Science and Engineering
Narbutta 85, 02 524 Warszawa
POLAND

phone: (48 22) 660 84 46
fax: (48 22) 48 48 75





From: Henry Eichelberger :      heichelb-at-binghamton.edu
Date: Thu, 9 Dec 1999 12:34:15 -0500
Subject: RE: Filters for List Serve?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


One List yes. But might I suggest using a * or # symbol preface. Thus
prefacing the Subject with a *B/ for biological interest, *M/ for Materials.
No prefacing of the Subject would indicate universal interest. Most E-mail
filters should be able to recognize *B/, *M/ in the Subject heading and
filter the E-mail out or into separate folders or mailboxes. For those who
forget to use the */ preface or simply don't want to be bothered, the
message will be available to all.
Cheers,
Henry

Henry Eichelberger, Manager
Electron Microscopy Facility
Department of Biological Sciences
Binghamton University
Binghamton, NY 13902-6000

phone: (607) 777-2682
fax: (607) 777-6521
e-mail: heichelb-at-binghamton.edu






From: skmenon-at-nps.navy.mil
Date: Thu, 9 Dec 1999 09:40:21 -0700
Subject: RE: GEN: Electron microscope as a turning point in history?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Some articles on history of the electron microscope :

Physica Review Vol 58 (1940) p.57
J. Appl. Phys. vol. 14 (1943) p. 434
Physics Today vol 15 (1962) p. 106
British Journal of Appl. Phys. Vol 13 (1962) p. 197
Science Vol 142 (1963) p. 185

Also see the books :
Fundamentals of TEM : R.D. Heidenreich
Introduction to EM : C.E. Hall
The world of the EM : RWG Wyckoff

Hope this helps

sarath

Sarath K Menon
Research Associate Professor
Department of Mechanical Engineering
Naval Postgraduate School
Monterey, CA 93943

Ph. # (831)-656-2551
FAX # (831)-656-2238







From: w-chiou-at-nwu.edu (Wen-An Chiou)
Date: Thu, 9 Dec 1999 11:54:16 -0500
Subject: Position Open for Scanning Electron Microscopist

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Scanning Electron Microscopist
Northwestern University

The electron probe instrumentation center (EPIC) at Northwestern
University has an immediate opening for a scanning electron microscopist.
EPIC is a part of the world renowned materials research center (MRC) and
the department of Materials Science & Engineering at Northwestern.

The scanning electron microscopist would be in charge of all of EPIC SEMs
(Hitachi S570, FE SEM S4500 and VP SEM 3500N), their accessories (EDS,
EBSD/OIM, LHe stage; systems) and the Hitachi FB-2000A focused ion beam
(FIB) system. All microscopes in EPIC are under full service contract.
Thus, the duties include mainly training students/users, development of
specialized techniques and applications, minor maintenance, record keeping
and billing. A BS/MS degree in physical/biological sciences is required.
The person must have hands-on experience in all aspects of SEM: specimen
preparation, EDS, digital acquisition, processing etc. All levels of
experience will be considered. Compensation would commensurates with
experience and qualifications.

Send cover letter, resume and three references to:

Prof. Vinayak P. Dravid, Director EPIC
Materials Science & Engineering
Northwestern University, 3013A MLSB
Evanston, IL 60208
E-mail: v-dravid-at-nwu.edu
Fax: (847) 491-7820

http://epic.ms.nwu.edu/epic/index.htm

Northwestern University is an Affirmative Action/Equal Opportunity
Employer.

Hiring is contingent upon eligibility to work in the United States.

************************************************
Wen-An Chiou
Department of Material Science and Engineering
Northwestern University
2225 N. Campus Dr.
Evanston, IL 60091

Tel: (847)-491-7807; Fax: (847)-491-7820
E-Mail: w-chiou-at-nwu.edu

http://epic.ms.nwu.edu/epic/index.htm
*************************************************







From: Walck. Scott D. :      walck-at-ppg.com
Date: Thu, 9 Dec 1999 13:18:00 -0500
Subject: filters for Haskaris water chillers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does anyone know where I can get the filters for a Haskaris water chiller
where I can use an AMEX card?
-Scott


Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)







From: David A. Cantor :      dacantor-at-mit.edu
Date: Thu, 09 Dec 1999 14:22:38 -0500
Subject: phase contrast vs. light microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


what are the advantages of a phase contrast microscope over a
conventional light microscope and why would i use one to examine embryos






From: Soumitra Ghoshroy :      ghoshroy-at-nmsu.edu
Date: Thu, 09 Dec 1999 12:40:31 -0700
Subject: One list please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think one list is the way to go. I am a biological person, but am
learning a lot from the materials postings.





Soumitra Ghoshroy Ph.D.
Electron Microscopy Lab
Box 3EML
New Mexico State University
Las Cruces, NM 88003-8001
Tel: 505-646-1531/3600
Fax: 505-646-5665
e-mail: ghoshroy-at-nmsu.edu





From: Laszlo Veto :      vgraphic-at-bcinternet.net
Date: Thu, 09 Dec 1999 10:05:16 -0800
Subject: Re: Please, only one list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I do agree and support to maintain one list only. It is beneficial to all.

Laszlo J. Veto








From: Geoff McAuliffe :      mcauliff-at-UMDNJ.EDU
Date: Thu, 09 Dec 1999 15:11:35 -0500
Subject: Re: GEN: Electron microscope as a turning point in history?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


"pjrandsjr-at-worldnet.att.net"-at-sparc5.microscopy.com wrote:

} Colleagues... This is a broad enough question that it can have many
} answers. Would any of you like to help me answer this for this student?
}
} I would think answers to both the list and to this student would be useful
} since he is not a member of the Server.
}
} Nestor
} Your Friendly Neighborhood SysOp
} -----------------------------------------------------------------------
}
} Email: pjrandsjr-at-worldnet.att.net
} Name: Philip Randolph
} School: Gonzaga Preparatory School
}
} Question: We are looking into the development of the electron microscope as
} a turning point in history. What specific information can you provide that
} will give us a sense of the effect of the advent of the electron microscope
} on the cultural, social and economic environment of society? Thank you for
} your time in considering this question.

They might want to look at "Picture Control, The Electron Microscope and
the Transformation of Biology in America, 1940-1960", by Nicholas Rasmussen.
Stanford Univ. Press, 1997. It got a good review in Science a while ago.

Geoff
--
**********************************************
Geoff McAuliffe, Ph.D.
Neuroscience and Cell Biology
Robert Wood Johnson Medical School
675 Hoes Lane, Piscataway, NJ 08854
voice: (732)-235-4583; fax: -4029
mcauliff-at-umdnj.edu
**********************************************







From: A. Kent Christensen :      akc-at-umich.edu
Date: Thu, 09 Dec 1999 15:12:27 -0500
Subject: (BP) One list

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would also vote for one list. I would suggest that the subject should
begin with one or two letters in parentheses: (B) = Biological. (P) =
Physical. (BP) Of possible general interest to microscopists.

Regards, Kent

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
A. Kent Christensen, Professor Emeritus
Department of Cell and Developmental Biology, Medical Science II Bldg.
University of Michigan Medical School, Ann Arbor, MI 48109-0616
Office: Room 2703A Med. Sci. II Bldg.
E-mail: akc-at-umich.edu, Tel. (work) (734) 763-1287, Fax. (work) (734)
763-1166
http://www.umich.edu/~akc/
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~





From: Missy Josephson :      ejosephs-at-neuron.uchc.edu
Date: Thu, 09 Dec 1999 16:50:37 -0500
Subject: microscopy list archive (at USF?)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I used to have a bookmarked website for an archive of biologically related
messages to the list that I think was maintained by someone at U. of South
FL. The listings were organized by subject matter, which made the resource
very handy. Unfortunately, I believe the bookmark got lost during a recent
computer installation and software reload. If anyone has any information on
an archive like this, would you please send me the address?

Many thanks,
Missy Josephson


Eleanor Josephson, DVM, PhD
University of Connecticut Health Center
Department of Anatomy MC-3405
263 Farmington Ave.
Farmington, CT 06030-3405
Ph.(860)679-2463
Fax (860)679-1274
ejosephs-at-neuron.uchc.edu





From: Rebecca Ai :      rebecca.ai-at-onsemi.com
Date: Thu, 09 Dec 1999 15:08:32 +0000
Subject: Re: One list please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Put yourself on both lists.


Rebecca Ai
Chemical and Surface Analysis Lab
Technology Development
On Semiconductor







From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Thu, 09 Dec 1999 17:03:38 -0600
Subject: Re: microscopy list archive (at USF?)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is the address of the lab page. Tips and Tricks is the top option on
the left.

http://www.biotech.ufl.edu/~emcl/

At 04:50 PM 12/9/1999 -0500, you wrote:

} I used to have a bookmarked website for an archive of biologically related
} messages to the list that I think was maintained by someone at U. of South
} FL. The listings were organized by subject matter, which made the resource
} very handy. Unfortunately, I believe the bookmark got lost during a recent
} computer installation and software reload. If anyone has any information on
} an archive like this, would you please send me the address?
}
} Many thanks,
} Missy Josephson






From: Mel Dickson :      m.dickson-at-unsw.edu.au
Date: Fri, 10 Dec 1999 10:20:19 +1100
Subject: Re: GEN: Electron microscope as a turning point in history?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A good place to start is the book

"Picture Control - the Electron Microscope and the Transformation of
Biology in America 1940-1960"

By Nicolas Rasmussen

Stanford University Press

Stanford California, 1997

ISBN 0-8047-2837-2





From: Barbara Foster :      mme-at-map.com
Date: Thu, 09 Dec 1999 18:44:41 -0500
Subject: Re: phase contrast vs. light microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear David,

As the name implies, Phase Contrast Microscopy detects differences in phase
(refractive index and/or thickness) between components in a system. In
order for this type of microscopy to work, however, you need very small
phase differences: small enough so that the light interacting with the
sample is slowed down or put out of step with the surrounding background
light by only a quarter of a wavelength. This system works really well with
things like cells but I would have serious concerns with a very 3D specimen
like your embryo. Other techniques which detect phase GRADIENTS (DIC and
Hoffman Modulation Contrast) would work much better.

For a further discussion of how each of these techniques operates, might I
suggest "Optimizing Light Microscopy"? Details of the book are on our website.

Best regards,
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************
MME is America's first national consortium providing
customized on-site workshops in all areas of
microscopy, sample preparation, and image analysis.


At 02:22 PM 12/9/99 -0500, David A. Cantor wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: paul nolan :      neysabob-at-hotmail.com
Date: Thu, 9 Dec 1999 21:53:32 -0600
Subject: GEN: Electron microscope as a turning point in history?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Philips Analytical put out a special issue of Electron Optics Bulletin
entitled "The Contribution of Electron Microscopy to Society"
This issue was from the Proceedings of a symposium on the topic of the same
title in Eindhoven in 1987 (dated, I know).

If you can obtain a copy of this it might be a good starting point. I would
try contacting Philips

Cheers
Paul

______________________________________________________
Get Your Private, Free Email at http://www.hotmail.com







From: paul nolan :      neysabob-at-hotmail.com
Date: Thu, 9 Dec 1999 21:53:01 -0600
Subject: GEN: Electron microscope as a turning point in history?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Philips Analytical put out a special issue of Electron Optics Bulletin
entitled "The Contribution of Electron Microscopy to Society"
This issue was from the Proceedings of a symposium on the topic of the same
title in Eindhoven in 1987 (dated, I know).

If you can obtain a copy of this it might be a good starting point. I would
try contacting Philips

Cheers
Paul

______________________________________________________
Get Your Private, Free Email at http://www.hotmail.com







From: =?iso-8859-1?B?wLHBuLW1?= :      jdyun-at-hanma.kyungnam.ac.kr
Date: Fri, 10 Dec 1999 12:21:27 +0900
Subject: Any biologist or materal scientist on the field of a clam shell?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear members:

My university runs a Education Center for the Gifted in Science and I am
involved in a mentorship program of the school. I teach three middle school
boys and girls in my group. We decided to do a project of studying 'clam
shell' with various analytical tools including SEM, XRD, and XRF. We have
already collected information on the amount of clam and oyster shell
production in my country.

Now we need some information on the science of clam shell as a reference,
such as the structure, mechanism of formation, function, composition, or
anything.

Do any of you have been in the field of clam or clam shell?
Do any of you know good references on the clamshell?
Any information would be very helpful to me and youngsters.

Jondo Yun
Department of Inorganic Materials Engineering
Center for Instrumental Analysis
Kyungnam University
449 Weolyeong-dong, Masan, 631-701, Korea
82-551-249-2697 (tel)
82-551-248-5033 (fax)
jdyun-at-hanma.kyungnam.ac.kr







From: r.cross-at-ru.ac.za
Date: Fri, 10 Dec 1999 08:00:55 +0200
Subject: Re: phase contrast vs. light microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello David

} what are the advantages of a phase contrast microscope over a
} conventional light microscope and why would i use one to examine
} embryos

Very simply, without going into details of the principles of the
technique, because it allows you to view unstained specimens.

Regards






Robin H Cross
Director : EM Unit, Rhodes University, Grahamstown, South Africa
R.Cross-at-ru.ac.za
tel: +27 46 603 8168 - fax: +27 46 622 4377
http://www.ru.ac.za/emu/index.htm

** remember that ICEM-15 in 2002 is in Durban, South Africa **





From: Keith Ryan :      KPR-at-wpo.nerc.ac.uk
Date: Fri, 10 Dec 1999 11:37:00 +0000
Subject: TEM - Philips EM300 disposal

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Microscopy Land

We are probably removing a Philips EM 300 transmission electron
microscope early next year in a departmenal reorganisation. Until
this year it was maintained by Philips, and currently by a third
party. It runs as well as when it was installed in April 1971.

Any offers would naturally be welcome, failing that, if anyone wants
to remove it, they probably can. We need the room. Maybe someone
would like parts for spares? Are there any more still running out
there? This one is 28 years old. It runs well at 80 kV but we have
never used it routinely at 100 kV, although it passes service muster
at 100 kV. It has the high resolution and goniometer stages. There is
a rotating specimen holder for the latter. The mercury diffusion pump
was removed some years ago. We will regret the passing of this trusty
instrument!

Regards - Keith Ryan




_______________________________
Dr. Keith Ryan
Marine Biological Association of the UK
The Laboratory
Citadel Hill
Plymouth
Devon PL1 2PB
England

Tel. ++44 (0)1752 633249
Tel. ++44 (0)1752 633279
The 279 number has an answering machine

Fax ++44 (0)1752 633102

e-mail: kpr-at-wpo.nerc.ac.uk





From: Robert Santoianni :      Robert_Santoianni-at-emory.org
Date: Fri, 10 Dec 1999 07:25:18 -0500
Subject: Haskris Chiller Filters

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Scott,
Try Philips service at 1-800-432-1734. I have a 20+y/o Haskris and
Philips has put a new filter on it during a pm on my EM201.
Bob Santoianni
Emory University Hospital
Atlanta, GA





From: Robert Palmer :      rjpalmer-at-dir.nidcr.nih.gov
Date: Fri, 10 Dec 1999 07:59:16 -0500
Subject: vendors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Does this list have a policy (written or otherwise) about vendor postings?
I have read the "welcome" message I got way back in the dark ages, and it
seems those rules prohibit direct advertisements. Certain postings I've
seen would seem to violate that rule (in spirit at least) in that the
signature lines identify the sender as an employee of organization X (as
required by the guidelines) which provides service Y (the advert). Does
this not qualify as an advertisement? Without a doubt, it is free air time
for the vendors involved. In certain cases, the violation has been less
discrete - when posters have requested information on a product/service,
vendor X replies on-list even though the poster's address is there for a
discrete reply. I'm just interested in where the line would be drawn (if
there is one).


Robert J. Palmer Jr., Ph.D.
Natl Inst Dental Craniofacial Res - Natl Insts Health
Oral Infection and Immunity Branch
Bldg 30, Room 310
30 Convent Drive
Bethesda MD 20892
ph 301-496-2088
fax 301-402-0396





From: L. D. Marks :      ldm-at-apollo.numis.nwu.edu
Date: Fri, 10 Dec 1999 07:35:41 -0600 (CDT )
Subject: Re: Haskris Chiller Filters

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You may want to check Haskris itself at www.haskris.com,
we have found them to be good. It is also worth mentioning
that their coolers are so common that our University Physical
Plant knows how to service them.

++++++++++++++++++++++++++++++++++++++++++++++++
Laurence Marks
Department of Materials Science and Engineering
Northwestern University
fax: (847) 491-7820
mailto:l-marks-at-nwu.edu
http://www.numis.nwu.edu
++++++++++++++++++++++++++++++++++++++++++++++++







From: Nestor J. Zaluzec :      zaluzec-at-Sparc5.Microscopy.Com
Date: Fri, 10 Dec 1999 08:21:07 -0600
Subject: Administrivia: Vendors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues....

I do monitor (but not edit) the submissions. When I see
that a posting has crossed that grey line which
loosely defines (the spirit) of our rules of conduct then
I contact the person(s) involved. Since I do not always read every message
I sometimes miss some transgressions, but people are pretty good
about this and I get off-line messages when
things appear to be, shall we say, in violation of conduct.

As a general rule most vendors are VERY VERY good and do not
post advertising. I do get complaints from Company Y about Company X
and usually after review a solution is found. The most common problems
are experienced with "new" users who do not
yet have experience in this forum and a gentile
reminder is sufficient. There are also the grudge matches which
happen and then I have to play mediator a job which is no fun
at all. (Hmm... maybe I should ask for a raise?).

Anyway, to speak to the question which Bob Plamor raised,
the rules of conduct (look at the WWW SIte the URL is on
EVERY Email message) do permit vendors to
say they make a product which can solve a particuliar problem.
They are asked to be concise and to the point and then contact
the individual off line. Vendors are one of our resources in my mind. Yes
they do sell things and as long as the reply answers a question
which was posted on the server this is valuable information and
totally within the spirit of the server. Sometimes they do get
verbose and when I notice this happening I let them know privately.

All in all, the vendors are important to us. They read closely what
is going on and the discussions surely influence what they think
of the scientific market place and what people need/want to conduct
their research. They are not just simply someone to be put up with
they are our colleagues abeit with different driving forces.
Some do only have only commerical motivations,
but I'm sure you will agree that there are those that have a passion
for the science as well and are there also to help us.

You should also remember that this is not a closed list and occassionally
spam/junk mail gets through the filter. These are completely
off topic and very hard to manage unless we change to a completely
closed list. That is something which I would like to avoid as it
not only means additional work, but makes us less accessible.


Nestor
Your Friendly Neighborhood SysOp








From: Julian Smith III :      smithj-at-Winthrop.edu
Date: Fri, 10 Dec 1999 09:38:27 -0500 (EST)
Subject: TEM repair: Coolwell chiller diagrams?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The venerable coolwell S100 chiller attached to our TEM responded to a
recent move from one building to another by dying. Does anyone have
schematics, or even an operator's manual for this thing. The guys from
physical plant and I need a starting place.....
TIA
Julian

Julian P.S. Smith III
Dept. of Biology
Winthrop University
Rock Hill, SC 29733
803-323-2111 x6427 (vox)
803-323-3448 (fax)







From: Soumitra Ghoshroy :      ghoshroy-at-nmsu.edu
Date: Fri, 10 Dec 1999 08:40:10 -0700
Subject: Job opening(correction)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Re: ELECTRON MICROSCOPY SPECIALIST
(Materials Science)
New Mexico State University


The correct website in the job description posted yesterday will be the
following:

http://www.nmsu.edu/~personel/postings/professional/164842211.html


Soumitra Ghoshroy Ph.D.
Electron Microscope Laboratory
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Phone: 505-646-3600/1531
Fax: 505-646-5665
E-mail:ghoshroy-at-nmsu.edu





From: A. Greene :      ablue-at-io.com
Date: Friday, December 10, 1999 8:20 AM
Subject: vendors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Dr. Palmer, Jr,

Yours is a very insightful question. My question is,
are vendors bad guys? Would you prefer to not have easy access to
instrumentation and technology?

I have been a member of MSA for about 20 years and now, as a vendor, am
delighted to be a sustaining member. Vendors, do not, in my opinion get
any "free air time." We pay for everything we get.

Alex Greene
SCIENTIFIC INSTRUMENTATION SERVICES, INC.
PMB-499, 1807 West Slaughter Lane, Number 200
Austin, Texas 78748 Phone: 512/282-5507 FAX 512/280-0702

Quality Electron Microscope Repair

-----Original Message-----
} From: Robert Palmer {rjpalmer-at-dir.nidcr.nih.gov}
To: Microscopy-at-Sparc5.Microscopy.Com {Microscopy-at-Sparc5.Microscopy.Com}






From: jennifer taylor :      jtaylor-at-stevens-tech.edu
Date: Fri, 10 Dec 1999 11:11:22 -0500
Subject: one list please

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to vote for one list with one or two letters in
paratheses. i.e.(B), (P), or (BP).

Jennifer Taylor






From: Huggins, Bradley J :      HUGGINBJ-at-BP.com
Date: Fri, 10 Dec 1999 11:32:40 -0500
Subject: This Listserver works

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Sorry to extend this thread at all, but I just feel compelled to remind any
of those who wish to separate/segregate the list - It is your combined
presence on this listserver that make it so absolutely functional! Our
depth is tremendous. Sure I can choose to ignore *Bio messages, or I could
refrain from signing up for the Bio sector of the listserver, if we were to
separate. So then I could focus on the work at hand. Not troubled by the
problems of those "Bio" guys trying to do that impossible staining or
labeling... But, then how would I ever know that some "bio-microscopists"
was doing things that may work for me. Or worse yet, how would I know that
some "bio-microscopist" could use my help and my recipe for xyz to embed his
zyx.

Yes we may have our dysfunctional moments, like any good family unit must,
just to maintain good perspective. (by the way, This is one of those
dysfunctional moments.) Please do not consider diminishing this powerful
resource. And Please Stay Tuned - to the whole picture.
Brad Huggins






From: Walck. Scott D. :      walck-at-ppg.com
Date: Fri, 10 Dec 1999 10:53:28 -0500
Subject: RE: filters for Haskris water chillers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Thank you to all that replied. The contact information for Haskris is

80 W. Seegers Rd.
Arlington Heights, IL 6005
Ph: 847-956-6420
fax: 847-956-6595

The strainers are $13 ea and yes they do take AMEX cards. Minimum order is
$35.

-Scott

} -----Original Message-----
} From: Walck. Scott D. [mailto:gls4590-at-smtpgate.go.ppg.com]
} Sent: Thursday, December 09, 1999 1:18 PM
} To: 'Microscopy'
} Subject: filters for Haskris water chillers
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} --------------------------------------------------------------
} ---------.
}
}
} Does anyone know where I can get the filters for a Haskaris
} water chiller
} where I can use an AMEX card?
} -Scott
}
}
} Scott D. Walck, Ph.D.
} PPG Industries, Inc.
} Glass Technology Center
} Guys Run Rd. (packages)
} P. O. Box 11472 (letters)
} Pittsburgh, PA 15238-0472
}
} Walck-at-PPG.com
}
} (412) 820-8651 (office)
} (412) 820-8161 (fax)
}
}
}





From: Richard Shuman :      rshuman-at-micrion.com
Date: Fri, 10 Dec 1999 12:56:34 -0500
Subject: GEN: Electron microscope as a turning point in history?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



For the 50th meeting of EMSA in Boston (1992) Sterling Newberry wrote a
wonderful book entitled "EMSA and Its People - The First Fifty Years". It
carries the Library of Congress number 92-72571. Sterling, along with his
wife Mary Lou, spent hundreds of hours interviewing many of the Charter
Members of our organization. Most of these interviews were video taped by
Sterling and then compiled into an "oral history" by Maurice Dumais. Both
Sterling's book and Maurice's "oral history" may still be available from the
MSA business office or borrowed from an attendee of the '92 Boston meeting.
In addition, John Reisner's series of "Reflection" columns in many of the
MSA Bulletins talk about the early pioneers and their contributions.

/rich shuman

-----Original Message-----
} From: "pjrandsjr-at-worldnet.att.net"-at-Sparc5.Microscopy.Com
[mailto:"pjrandsjr-at-worldnet.att.net"-at-Sparc5.Microscopy.Com]
Sent: Thursday, December 09, 1999 9:06 AM
To: Microscopy-at-Sparc5.Microscopy.Com


Colleagues... This is a broad enough question that it can have many
answers. Would any of you like to help me answer this for this student?

I would think answers to both the list and to this student would be useful
since he is not a member of the Server.

Nestor
Your Friendly Neighborhood SysOp
-----------------------------------------------------------------------

Email: pjrandsjr-at-worldnet.att.net
Name: Philip Randolph
School: Gonzaga Preparatory School


Question: We are looking into the development of the electron microscope as
a turning point in history. What specific information can you provide that
will give us a sense of the effect of the advent of the electron microscope
on the cultural, social and economic environment of society? Thank you for
your time in considering this question.










From: Michael Bode :      mb-at-soft-imaging.com
Date: Fri, 10 Dec 1999 10:47:26 -0700
Subject: RE: vendors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is gonna be interesting....

As one of those vendors (and a former 'non-vendor') on the list, I would
like to say a few words to this. I am not speaking for any other vendor
on the list, this is my own opinion.

I think, the entire issue of vendors of this list is:

a) a gray area and will always stay that way
b) something that is handled very well by Nestor.

Personally, when I read a listing I feel the urge to respond to, I make
the following decision: Does the posting I intend to put on the list
include information that may interest more people than the person whose
message I am replying to?

If the answer is Yes, I go to the list.
If the answer is No, I respond through email.

An example:

If somebody posts: "Are there ways to reduce noise in digital TEM
images?", I'd post on the list. If the question is: "Why doesn't
function X work with software Y on my TEM Z", I'd stick with email. Of
course these are judgment calls and some people may have broader or
narrower definitions for putting something on the list.

I don't see the point, however, to hide my affiliation. I work for a
company and not for a University, but why would that disqualify me from
making contributions to the list? If you are concerned about more
blatant misuse of the list, I don't think there are many, and I am sure
that Nestor would deal with those. On the other hand I might mention my
company or one of its products in the posting, but there is not much I
can do about it: I cannot speak for other vendors so I have to stick to
what I can use, which is my own products. In the example above: I can
use examples from our own software, but not from any other as I don't
want to misrepresent the other software.

One thing that many vendors put in are disclaimers at the end (I must
admit that I have forgotten a few times) to make sure that everybody
knows that they are interested in selling something. Whether that is
useful I don't know. I guess one could filter for "disclaimer" and throw
those out. On the other hand exactly that could be interpreted as an
advertising.

But what, for example, about conference announcements. Conferences are a
business and if you exclude all commercial listings, you would have to
exclude them as well. I think, that would be disservice to the
community.

In a nutshell: I don't think this issue is a Yes/No issue but rather an
issue of balance between usefulness and abuse. Of course I am biased,
but I think Nestor (or perhaps the vendors?) do a good job at keeping
this ratio at a good level.

but then again... this is only my opinion. (Does this posting count as
covert advertising???)

Michael Bode, Ph.D.
Soft Imaging System Corp.
1675 Carr St., #105N
Lakewood, CO 80215
===================================
phone: (888) FIND SIS
(303) 234-9270
fax: (303) 234-9271
email: mailto:info-at-soft-imaging.com
web: http://www.soft-imaging.com
===================================

*******************************************************
Disclaimer:
Soft Imaging System produces and sells image acquisition
and processing systems. We therefore have a vested
interest in some of the items mentioned above.
*******************************************************

} ----------
} From: Robert Palmer[SMTP:RJPALMER-at-DIR.NIDCR.NIH.GOV]
} Sent: Friday, December 10, 1999 5:59:16 AM
} To: Microscopy-at-sparc5.microscopy.com
} Subject: vendors
} Auto forwarded by a Rule
}
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Does this list have a policy (written or otherwise) about vendor
postings?
I have read the "welcome" message I got way back in the dark ages, and
it
seems those rules prohibit direct advertisements. Certain postings I've
seen would seem to violate that rule (in spirit at least) in that the
signature lines identify the sender as an employee of organization X (as
required by the guidelines) which provides service Y (the advert). Does
this not qualify as an advertisement? Without a doubt, it is free air
time
for the vendors involved. In certain cases, the violation has been less
discrete - when posters have requested information on a
product/service,
vendor X replies on-list even though the poster's address is there for a
discrete reply. I'm just interested in where the line would be drawn
(if
there is one).


Robert J. Palmer Jr., Ph.D.
Natl Inst Dental Craniofacial Res - Natl Insts Health
Oral Infection and Immunity Branch
Bldg 30, Room 310
30 Convent Drive
Bethesda MD 20892
ph 301-496-2088
fax 301-402-0396





From: Gordon Vrololjak :      gvrdolja-at-nature.Berkeley.EDU
Date: Fri, 10 Dec 1999 10:58:17 -0800 (PST)
Subject: Re: Any biologist or materal scientist on the field of a clam shell?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello,
I have seen a talk by P. K. Hansma on the study they did of abalone
shells.

To paraphrase him:
Seashells are interesting as they are 97% CaCO3, 3% organic material; and
are 3000x more fracture resistant than a single crystal of CaCO3. =20
Industry is currently doing research on using seashell like material to
replace plastics and wood since it would be fire resistant and non-toxic. =
=20
Since seashells grow slowly, industry needs to grow them faster.

Studies by SEM showed abalone mother of pearl to consist of stacks of
plates of CaCO3. (excuse my poor diagram)

_____ ______ ______ ___________
| | | | | | | |
----- ------ ------ -----------
-- ------ ------ ------- ------
| | | | | | | | |
-- ------ ------ ------- ------
----- ------ --------- -----
| | | | | | |
----- ------ --------- -----

Crack energy is dissipated by the layers stacked randomly. Composites of
CaCO3 materials manufactured similar to above didn't work. The reason for
the strength of the abalone shell was found to be due to a protein binding
between the tablets (Lustrin A). Immunogold labeled SEM showed the
material to appear as strands between plates: (again, excuse my diagram)

-----------
\/ \/
| |=20
| |
| |
| |
/\ /\
------------

Pulling CaCO3 plates apart in the seashell causes the protein structure to
unfold and many intermediate strength protecting bonds are broken
slowly. Relaxing the force causes the protein bonds to reform.

Some useful papers may be:
Zaremba, CM; Morse, DE; Mann, S; Hansma, PK; and others.
Aragonite-hydroxyapatite conversion in gastropod (abalone) nacre.
CHEMISTRY OF MATERIALS, 1998 DEC, V10 N12:3813-3824.

10. Schaffer, TE; IonescuZanetti, C; Proksch, R; Fritz, M; and others.
Does abalone nacre form heteroepitaxial nucleation or by growth
through mineral bridges? (vol 9, pg 1731, 1997)
CHEMISTRY OF MATERIALS, 1998 MAR, V10 N3:946-946.
Pub type: Correction/Addition.

12. Shen, XY; Belcher, AM; Hansma, PK; Stucky, GD; and others.
Molecular cloning and characterization of lustrin A, a matrix
protein from shell and pearl nacre of Haliotis rufescens.
JOURNAL OF BIOLOGICAL CHEMISTRY, 1997 DEC 19, V272 N51:32472-32481.

18. Walters, DA; Smith, BL; Belcher, AM; Paloczi, GT; and others.
Modification of calcite crystal growth by abalone shell proteins: An
atomic force microscope study. BIOPHYSICAL JOURNAL, 1997 MAR, V72 N3:14=
25-1433.

26. ZAREMBA CM; BELCHER AM; FRITZ M; LI YL; and others.
CRITICAL TRANSITIONS IN THE BIOFABRICATION OF ABALONE SHELLS AND
FLAT PEARLS. CHEMISTRY OF MATERIALS, 1996 MAR, V8 N3:679-690.
=20
1. Shen, XY; Belcher, AM; Hansma, PK; Stucky, GD; and others.
Molecular cloning and characterization of lustrin A, a matrix protein=
=20
from shell and pearl nacre of Haliotis rufescens.
JOURNAL OF BIOLOGICAL CHEMISTRY, 1997 DEC 19, V272 N51:32472-32481.

22. Taylor, John David.
The shell structure and mineralogy of the Bivalvia, by John David
Taylor, William James Kennedy [and] Anthony Hall. London, British Museum
(Natural History), 1969-73. Series title: Bulletin of the British Museum
(Natural History). Zoology. Supplement ; 3. Series title: British Museum
(Natural History) Bulletin. Zoology, v. 22, no. 9.

I hope this information may be of some use to you. The formation of
seashells is very interesting chemically since it involves chemical
precipitation aided by an organic biological template. Often crystal
forms predicted to be unstable thermodynamically are precipitated due to
the effect of the organism and the biological material it produces.

I am interested in researching sea shells myself, but haven't had the
opportunity to work with anyone. I have acess to TEM, SEM, and AFM
instruments, so if someone is interested in working together, drop me a
line.=20

\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\/\=
/\
Gordon Ante Vrdoljak 1 Cyclotron Road
ICQ 23243541 http://nature.berkeley.edu/~gvrdolja MS90-1116
GAVrdoljak-at-lbl.gov Ernest Orlando
phone (510) 495-2829 Lawrence Berkeley
fax (510) 486-7797 National Laboratory
cell (510) 290-6793 Berkeley CA 94720

On Fri, 10 Dec 1999, =C0=B1=C1=B8=B5=B5 wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=
=20
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
} =20
} =20
} Dear members:
} =20
} My university runs a Education Center for the Gifted in Science and I =
am
} involved in a mentorship program of the school. I teach three middle sch=
ool
} boys and girls in my group. We decided to do a project of studying 'clam
} shell' with various analytical tools including SEM, XRD, and XRF. We have
} already collected information on the amount of clam and oyster shell
} production in my country.
} =20
} Now we need some information on the science of clam shell as a referenc=
e,
} such as the structure, mechanism of formation, function, composition, or
} anything.
} =20
} Do any of you have been in the field of clam or clam shell?
} Do any of you know good references on the clamshell?
} Any information would be very helpful to me and youngsters.
} =20
} Jondo Yun
} Department of Inorganic Materials Engineering
} Center for Instrumental Analysis
} Kyungnam University
} 449 Weolyeong-dong, Masan, 631-701, Korea
} 82-551-249-2697 (tel)
} 82-551-248-5033 (fax)
} jdyun-at-hanma.kyungnam.ac.kr
} =20
} =20
} =20
} =20







From: SEMTRADER-at-aol.com
Date: Fri, 10 Dec 1999 14:33:02 EST
Subject: The List

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


It appears as though there hasn=92t been much contribution to the list in th=
is thread.

Let get back to microscopy.! =20

If you would like to offer your commercial products post it to
http://www.msa.microscopy.com/News/NewsListings.html

My 2 Cents
Michael A.






From: Barbara Foster :      mme-at-map.com
Date: Fri, 10 Dec 1999 15:02:02 -0500
Subject: Re: Any biologist or materal scientist on the field of a clam

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Yondo,

I notice that you have not included light microscopy in your list. We
typically recommend, especially in this sort of case, that your students
start with a simple hand lens, then move to a stereo microscope ... both of
which will give a lot of information about context and macro structures.
They should then move to the compound microscope, looking at both the inner
and outer surface of the shell. Polarized light will give some amazing
information about the orientation of the calcium carbonate and how it was
layed down when the shell was formed. None of this information is
available with the other techniques cited.. and I wouldn't want either you
or your students to miss out on this part of the "mystery".

Good hunting and best regards,

Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************
MME is America's first national consortium providing
customized on-site workshops in all areas of
microscopy, sample preparation, and image analysis.


At 12:21 PM 12/10/99 +0900, =C0=B1=C1=B8=B5=B5W1 wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20
school
} boys and girls in my group. We decided to do a project of studying 'clam
} shell' with various analytical tools including SEM, XRD, and XRF. We have
} already collected information on the amount of clam and oyster shell
} production in my country.
}
} Now we need some information on the science of clam shell as a reference,
} such as the structure, mechanism of formation, function, composition, or
} anything.
}
} Do any of you have been in the field of clam or clam shell?
} Do any of you know good references on the clamshell?
} Any information would be very helpful to me and youngsters.
}
} Jondo Yun
} Department of Inorganic Materials Engineering
} Center for Instrumental Analysis
} Kyungnam University
} 449 Weolyeong-dong, Masan, 631-701, Korea
} 82-551-249-2697 (tel)
} 82-551-248-5033 (fax)
} jdyun-at-hanma.kyungnam.ac.kr
}
}
}
}
}





From: O'Neil, David :      David.O'Neil-at-nrc.ca
Date: Fri, 10 Dec 1999 16:16:59 -0400
Subject: EM as a turning point in history

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Just another interesting note on EM in history. The Arts & Entertainment
channel on the tube did a "100 greatest achievements of the Century" and
Electron Microscopy was #59 (or somewhere close to that). Right up there
with tungsten lighting and insulin. Perhaps the student could check with
the producers of that program.

} } } } } Keep it to one list { { { { { { { { {

David O'Neil
National Research Council of Canada
Institute for Marine Biosciences
1411 Oxford Street
Halifax, NS B3H 3Z1
ph: (902)426-8258
fax: (902)426-9413
e-mail: david.o'neil-at-nrc.ca





From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 10 Dec 99 17:16:54 -0500
Subject: Bio-zebra fish

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Hi,
I have a researcher who plans to embed developing zebra fish to study
development of the inner ear. Samples will be fixed and embedded up to
age 30 days. We are not concerned about early stages but did not know when
calcification starts and if this would be a problem for the late stages.

Has anyone had experience with fish or other embryos at the early
stages of calcification? Did you need to do any decalcification and, if so,
what procedure did you use? We are not interested in the bones themselves
but the soft tissue and are concerned that calcified skull could cause
tearing of the nearby tissues when the block is sectioned. Also any special
protocols for optinmum fixation?

Thanks in advance for your help.
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057






From: Maria Lucia Ribeiro Caldas :      caldasml-at-amcham.com.br
Date: Fri, 10 Dec 1999 21:20:08 -0200
Subject: Chromotrope aniline blue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by amhost4.amcham.com.br (8.8.6/8.8.6) with ESMTP id VAA24102
for {Microscopy-at-sparc5.microscopy.com} ; Fri, 10 Dec 1999 21:15:16 -0200 (EDT)
Message-ID: {38518AA5.EF6BEE29-at-amcham.com.br}


Dear all

I need to do special staining for giant mitochondria in transplant
kidney bx, it's chromotrope aniline blue. The ref. I found was
incomplete. I will be very happy if anyone coul send me the formula. If
you also have the formula of Sirius red for interstitial fibrosis, my
happiness will be complete.

Thanks in advance

Lucia Caldas
Rua Pres. Backer 234/604 BII
Niter=F3i Icara=ED RJ
Brasil 24230-041
Cell phone: 55-021-99587768.






From: Vachik Hacopian :      vhacopian-at-wellesley.edu
Date: Fri, 10 Dec 1999 19:21:55 -0400
Subject: to Haskris chiller/recirculator owners

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have had an R75 and an R100 for many years now. I bought them based on
Philips' recommendation. We have had to replace the water pump on both
units about once a year. According to Haskris, that's too frequent. What
happens is that the brass coupling between the pump and the motor wears
unevenly. We run it until the pump fails. Haskris suggested installing a
new coupling every 6 months as routine p.m. There has been one occasion
where the pump has failed in 5 months. The coupling was worn down to 20%
of its mass. Has anyone else experienced this?

Vachik Hacopian







From: Paul Webster :      pwebster-at-mailhouse.hei.org
Date: 10 Dec 99 17:17:36 -0800
Subject: Re: vendors

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Dear All,

Give these guys a break! Where would we be without vendors?

Not only are they the source of ALL our equipment but in many instances =
are also a very valuable source of information. They know the equipment =
better than most, can identify trends faster, do visit all the labs =
working with microscopes so can see new technologies before we do, and can =
provide solutions to problems that may seem impossible when viewed up =
close. =

I think the moderation of the list is tough enough, both from the =
contributors and from Nestor, to ensure that what gets through is tasteful =
and concise. I encourage vendors to continue contributing not only =
because they are at the front of what is happening, but also because many =
are established and experienced microscopists too. When your reps come =
through again, try talking with them. You might learn a thing or two.


Regards, =

Paul Webster, Ph.D
House Ear Institute
2100 West Third Street
Los Angeles, CA 90057
phone:213 273 8026
fax: 213 413 6739
e-mail: pwebster-at-hei.org
http://www.hei.org/htm/aemi.htm

PS. On the subject of splitting the list - As a biologist I think we have =
lots to learn from the materials people, especially when it comes to =
sectioning. Having talked with material scientists, I think there are one =
or two things we can pass on too. Sorry Lorry, but my vote is to stick =
with one list.

Finally: A note to Nestor - I like the idea of you giving out "gentile =
reminders" but I need to know if jewish ones are more or less tough?






From: Christian Kuebel :      ckuebel-at-engin.umich.edu
Date: Fri, 10 Dec 1999 21:01:41 -0500
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by srvr5.engin.umich.edu (8.9.1a/8.9.1) with SMTP id UAA17637
for {microscopy-at-microscopy.com} ; Fri, 10 Dec 1999 20:55:37 -0500 (EST)
Message-Id: {3.0.5.32.19991210210141.0098e440-at-srvr5.engin.umich.edu}
X-Sender: ckuebel-at-srvr5.engin.umich.edu
X-Mailer: QUALCOMM Windows Eudora Light Version 3.0.5 (32)


I would appriciate it, if you would subscribe me to the microscopy list sever.

Thanks,

Christian Kuebel


-------------------------------------------------
Dr. Christian Kuebel
University of Michigan
Department of Materials Science and Engineering
2125 H. H. Dow Building
2300 Hayward Street
Ann Arbor, MI 48109-2136
Phone: (734) 763-4196





From: Gordon Couger :      gcouger-at-rfdata.net
Date: Sat, 11 Dec 1999 02:37:55 -0600
Subject: Nachet help

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I purchased a Nachet 300 tricoular with what appears to be a phase
condenser, G.C. 10x eyepieces a P8x in the camera tube, oculars: 60 1.00, 40
.65, 20 .55 and 10 .27. The thing that I don't understand is there are two
dove tail sliders abover the oculars that can be inserted or removed from
the light path and one has a very fine adjustment in the light path.

I would like a manual in English but I doubt one is available. I will go by
the library and find a book on phase scopes to learn the adjustememt
procecure.

It seems like a well built scope except the 35 mm camera that came with it
looks like it came from a box of cracker jacks. I know that all the camera
is is a box that holds film but this is the a sorry excuese for a 35 mm
back. That is not much of a concern as I will be replacing it with a video
camera.

Can some on tell me what the sliders are above the objectives. If it were a
poloriszed scope they would be phase plates. They don't give enough color to
be phase plates unless the polorizer is missing.

Thanks
Gordon

Gordon Couger gcouger-at-couger.com

Stillwater, OK www.couger.com/gcouger
405 624-2855 GMT -6:00 www.couger.com/gcouger







From: Debby Sherman :      sherman-at-btny.purdue.edu
Date: 11 Dec 99 09:56:03 -0500
Subject: Re: to Haskris water chiller owners

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


We have had one Haskris R100 with two pumping units for about 30 years.
We just replaced the pump on one unit which, I believe, is only the second
time it has been replaced. The other one has also been replaced at least
once as well. The original motors were on the unit although the
compressor was replaced once and the copper piping has been replaced/repaired over
the years. Another double pump R100 unit has had similar types and
frequency of repairs.

Our compressors have gone out primarily beause they are water cooled
and were on hard water. Mineral build-up restricted the cooling water
causing failure. Now our maintenance crew runs dilute acid through them
yearly to prevent this problem.All work is done by our in-house plumbing and
refrigeration crew.
Debby

Debby Sherman, Manager Phone: 765-494-6666
Microscopy Center in Agriculture FAX: 765-494-5896
Dept. of Botany & Plant Pathology E-mail: sherman-at-btny.purdue.edu
Purdue University
1057 Whistler Building
West Lafayette, IN 47907-1057

Vachik Hacopian wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America






From: Rosemary Walsh :      rw9-at-psu.edu
Date: Sat, 11 Dec 1999 10:24:48 -0500
Subject: Re: Any biologist or materal scientist on the field of a clam

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Gordon,
Thank you for your thorough posting especially the diagrams and references
you included.

!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!one list!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!keep
vendors' input!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
Rosemary Walsh






From: rfelten-at-Macdermid.com
Date: Sat, 11 Dec 1999 12:49:40 -0500
Subject: Opinions on Steromicroscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Rick Felten-at-MACDERMID
12/11/99 12:49 PM
I specialise in using the SEM to characterize bulk surfaces, but as a
prelude to the analysis of each sample I always preview it w/ a Nikon low
mag stereo inspection microscope. This light microscope often gives me the
info I seek, and I simple use the SEM to confirm or prove my expectations.
I am not knowledgeable about light microscopy, but when I compare the
information I get from my Nikon (20-30 years old) to many other
stereoscopes, I find that all the others at my company give very poor
performance. I want to purchase a stereo microscope for my kids to grow up
with. Something that matches our visual world but simply magnified To me
this means excellent resolution, brightness, and 3-D. I was wondering who
are the "undisputed" quality vendors and who are the vendors that produce a
fine product, but don't have a name like Carl Zeise, Nikon and Leica. How
much does one have to spend to get the quality that I am used to in my
Nikon. How high in magnification can one achieve and still have good
stereo information? I am interested in only reflectance mode work. Has
anyone recently purchased a good basic stereo microscope that impressed
them. I would like to have the ability to ditigize at a later date, so I
assume I need a phototube? I see a ad in Cole palmer for a Meiji
microscope about $2000. I have also seen prices that ranged from $300 to
$7000, where is a good cut off for price/performance ratio?
Ric







From: Marek Malecki :      mmm-at-biomail.ucsd.edu
Date: Sat, 11 Dec 1999 17:44:07 -0600
Subject: piezo-focus

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Would anyone be willing to share experience with piezo-focus devices from
various manufacturers?

Marek Malecki, M.D., Ph.D.
Director and Principal Investigator

Molecular Imaging Laboratories
University of California at San Diego

address: 1500 Suite Bonner Hall, 9500 Gilman Drive, La Jolla, CA 92093-0368
phone: 8588223373
fax: 8588223715
pager: 8586161420
e.mail: mmm-at-ucsd.edu
e.pager: 1620024619-at-alphapage.airtouch.com
www site: http://mil.ucsd.edu/
ftp site: mil.ucsd.edu







From: bobrob-at-uswest.net
Date: Sat, 11 Dec 1999 18:05:43 -0700
Subject: TEM: Chiller/Recirculator Wanted

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have an immediate need to locate and purchase a suitable
water chiller/recirculator in good used condition. The unit must
be an air-to-air-type with a cooling capacity of approx 15,350 BTU
per hour. Likely models include Haskris R100 or Neslab HX 150.
Would prefer a unit geographically nearby.

Please respond directly to:

Bob Roberts
EM Lab Services, Inc.
2409 S. Rural Rd
Tempe, Arizona 85282
(480) 967-3946







From: earlw-at-pacbell.net
Date: Sat, 11 Dec 1999 23:29:27 -0800
Subject: Re: JEOL JSM-35CF Spectrometer & Optical Microscope

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,

The above equipment has been spoken for. Thanks to all who responded.


Earl Weltmer






From: =?iso-8859-1?Q?Varga_L=E1szl=F3?= :      varguc-at-freemail.c3.hu
Date: Sun, 12 Dec 1999 13:03:22 +0100
Subject: Info about Microscan 9

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This subject comes up now and then, and please allow me to add my
commercially oriented view. The original 'Internet' had usage policies
that forbade the conveyance of commercial 'advertising', although many of
the participants were large commercial concerns. Nearly a decade ago, the
Internet was commercialized, leading to the present situation, such as it
is, where the usage rules simply don't apply. This was a result of the
turnover of the Internet to commercial, rather than governmental control.

Obviously, there are blatant abusers out there in the nasty world of the
Internet. IMHO, one of the worst is AOL, the leading ISP. However, there
are also relative bastions of calm such as this list. Primarily due to its
esoteric nature, and the efforts of Nestor, this list has remained
relatively free of blatant commercialism.

I have never started a thread, but have often responded to one. There have
been one or two occasions over the years where our friendly sysop (Nestor)
has contacted me regarding the suitability of a posting I made. In each
case, I believe we have managed to come to a reasonable mindset where I
have not been overtly commercial and the list has not been harmed by my
postings.

The bottom line is that academic purity is not damaged by the honest
presentation of commercial concerns, and commercial purity need not be
compromised by academic concerns. On the contrary, a substantive admission
of orientation, both on the commercial and academic sides, can only enhance
the interchange.

The subject of signature lines is simply specious. A proper signature line
merely identifies one as belonging to one organization or another, and, as
such, helps the viewer to determine one's affiliations and potential
biases. Whether a university, a company or a high school might benefit
from a particular signature line is really up the the viewer. If you think
that you should be the arbiter of what each viewer reads in those lines,
then I suggest that you might consider therapy.

I believe that most viewers, including myself, can come to reasonable
conclusions based on the affiliations presented. This is far preferable to
a list where no affiliations are presented and one is left not knowing the
source of a particular view. No doubt in your view, this represents an
unreasonable black vs. white approach, but the truth is that you either
allow affiliations or you don't. The only alternative is to allow
affiliations only for a chosen few and that doesn't cut it in the modern
world.

In regards to your statement that "Without a doubt, it is free air time for
the vendors involved", how many academic postings are made with
affiliations noted in the signature line with the intention of aggrandizing
an individual, department or organization? Let's be honest, a signature
line, whether private or public, is an attempt to tie an individual to an
organization that may carry some weight to that individual's view or enh
ance the organization's stature should that individual's view be accepted.
Academic ventures are surely not immune to financial concerns.

The purpose that all of us should follow is the enhancement of the
scientific principles that through the centuries have provided the
enlightenment we have pursued. Rather than getting caught up in the
micro-management of that process, we should be involved in the active
management and stewardship of the principals that have brought us so far.


Allen R. Sampson
Advanced Research Systems
317 North 4th. Street
St. Charles, IL 60174
PH 630.513.7093 FAX 630.513.7092 Email: ars-at-sem.com

-----Original Message-----
} From: Robert Palmer [SMTP:rjpalmer-at-dir.nidcr.nih.gov]
Sent: Friday, December 10, 1999 6:59 AM
To: Microscopy-at-sparc5.microscopy.com


Dear list members,

I may get involved in the installation of a Cambridge Microscan 9 EPMA
donated to an institute in Hungary. I would highly appreciate any kind of
information about that model.

Laszlo Varga





From: Clarence A. Miller :      camill-at-rice.edu
Date: Sun, 12 Dec 1999 14:51:04 -0600
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please remove my name from the list.

Clarence A. Miller
Department of Chemical Engineering
Rice University
6100 S. Main Street
Houston, Texas 77005-1892
Telephone: 713-527-4904
FAX: 713-285-5478







From: Claudia Hayward-Costa :      LS_S562-at-crystal.kingston.ac.uk
Date: Mon, 13 Dec 1999 11:15:29 +0100
Subject: B: BEEM capsulas,embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Experts,

I am working with leukemia cells in suspension and spin them
down in conical BEEM capsules prior to fixation. All subsequent
steps (osmication, dehydration, infiltration with epon 812) are
performed in these tubes.
Unfortunately the blocks are very difficult to cut. The tips are with
holes and inside the blocks there are "craters" and spongy areas
that make it very difficult to cut proper sections.
The pellets are minute, therefore I like the convenience of the
conical tubes but I have to cut "big" chunks away with the razor
blade before I get reasonable sections which makes it a quite
wasteful job.
The absolute EtOH is dried over Cu2SO4 in a dialysis tube and the
cells are in filtrated twice within 24 hours.

I know that one way out of this dilemma would be to handle the
pellet as a tissue - but the experiment is so time consuming that I
still do not want to risk having a crumbling pellet after 2 days work.

Thanks in advance for your kind help.

Claudia


Dr. C. Hayward-Costa
School of Life Sciences
Kingston University
Penrhyn Road, Kingston upon Thames
Surrey KT1 2EE, UK
++44(0)181 547 2000 x 2240
Email: c.hayward-at-kingston.ac.uk





From: Herczeg, Laszlo :      lherczeg-at-edax.com
Date: Mon, 13 Dec 1999 07:12:59 -0500
Subject: Please remove my name from the list.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Please remove my name from the list.


Laszlo I. Herczeg, Sr. Systems Engineer
EDAX Inc.
91 McKee Drive Mahwah, NJ 07430
201-529-6224
Email: LHerczeg-at-edax.com






From: John Catino :      jwcatino-at-concentric.net
Date: Mon, 13 Dec 1999 07:54:39 -0600
Subject: where I can get cryo-SEM done

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



}
} I recommend Dr. Cheng Huang {chuang-at-ccs.carleton.ca} at Carleton University,
} Ottawa, Canada.
}
} John Catino
}
}
}
}
} "Tian_Huang-at-gillette.com"-at-sparc5.microscopy.com wrote:
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} } Hi, there,
} } Could anybody tell me where I can get cryo-SEM done? Thanks!
} }
} }
} } Tian
}







From: Bob Lippert :      blpprt-at-clemson.edu
Date: Mon, 13 Dec 1999 09:34:36 +0000
Subject: Image needed

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
I use a PowerPoint presentation to teach farmers, county extension
agents and horticulturalists about basic soil science.
I need a scanning electron microscope image of kaolinite. Only one
or two very good images are all that I need. Can anyone help with
this? I thought there might be a national library of SEM images for
public access but there doesn't seem to be one.
If you can help, please email me privately at:
blpprt-at-clemson.edu
Thanks!
Bob Lippert



*********************************************************
Dr. Bob Lippert
Dept. Crop and Soil Environmental Science
277 Poole Agricultural Center
Box 340359
Clemson University
Clemson, SC 29634-0359
Phone 864-656-3502
FAX 864-656-3443
EMAIL blpprt-at-clemson.edu
WEB SITE: http://hubcap.clemson.edu/~blpprt/bobweb/bobweb.html





From: William McManus :      billemac-at-biology.usu.edu
Date: Mon, 13 Dec 1999 11:26:20 -0700
Subject: Job opening(correction)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Out of curiosity, could you divulge the salary range for this position?

William McManus
Supervisor
Electron Microscopy Facility
Department of Biology
Utah State University
Logan UT 84322-5305

billEMac-at-cc.usu.edu
435-797-1920


-----Original Message-----
} From: Soumitra Ghoshroy [mailto:ghoshroy-at-nmsu.edu]
Sent: Friday, December 10, 1999 8:40 AM
To: Microscopy-at-Sparc5.Microscopy.Com


Re: ELECTRON MICROSCOPY SPECIALIST
(Materials Science)
New Mexico State University


The correct website in the job description posted yesterday will be the
following:

http://www.nmsu.edu/~personel/postings/professional/164842211.html


Soumitra Ghoshroy Ph.D.
Electron Microscope Laboratory
Box 3EML
New Mexico State University
Las Cruces, NM 88003
Phone: 505-646-3600/1531
Fax: 505-646-5665
E-mail:ghoshroy-at-nmsu.edu





From: John Grazul :      grazul-at-physics.bell-labs.com
Date: Mon, 13 Dec 1999 14:34:07 -0500
Subject: Need an EM lab designer in NYC area

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Listees,

We are in need of an architect to help us design a TEM room in a stand
alone structure. The building will be an empty shell, so we can do
pretty much as we please. The FETEM is used for high res work which
means it has some very special needs. If you know of a firm in this
area {NY, NJ, PA, CN} which can handle this please contact me via
e-mail.

Thanks,

John Grazul
Lucent Technologies
Murray Hill, NJ





From: Susan Belfry :      belfry-at-unb.ca
Date: Mon, 13 Dec 1999 16:06:21 -0400
Subject: LM: JB-4 resin and fish eggs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



I have no experience working with JB-4 resin and a graduate student is
trying to embed whole fish (trout) eggs for a histological examination.
The eggs are about 5 mm in diameter. So far, penetration into the yolk is
incomplete or not at all. Vacuum infiltration is not recommended because it
may speed up polymerization of the JB-4 resin and the student would like to
have the eggs intact during preparation. Does anyone have any suggestions?
Or can one point us to a literature source for this type of work?

Many thanks, Susan
**************************************
Susan Belfry
Electron Microscopy Unit
University of New Brunswick,
Bag Service #45111, Fredericton,
New Brunswick, E3B 6E1, Canada
Phone: 506-453-4887 Fax: 506-453-3583
http://www.unb.ca/emunit/
*************************************






From: Jo Dee :      jofish-at-burnham-inst.org
Date: Mon, 13 Dec 1999 12:29:42 -0800
Subject: Re: B: BEEM capsulas,embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Claudia,
I have processed many cell pellets in the 0.6ml microcentrifuge tubes (I
have never tried to use the BEEM capsules for processing). I have had no
problems processing and embedding the cells all in the same tube. It really
sounds like the cells are not thoroughly dehydrated. What kind of
dehydration series are you using? I have found that if I use a rinse with
100% acetonitrile after the 100% EtOH and then infiltrate with a 2:1, 1:1,
1:2 acetronitrile:eponate 12 series I have absolutely no problems with the
finished block. Also, before putting the pure eponate 12 resin in the tube
(the final step before polymerization) I remove every last bit of
acetonitrile with a small absorbent cotton swab. This way I am satisfied
that there is absolutely no water or other liquid around my cells which may
prevent the resin from getting all the way down into the tip of the tube.
I hope this works for you!
Good luck,
Jo Dee

Jo Dee Fish
Electron Microscopy Assistant
The Burnham Institute
10901 N. Torrey Pines Rd.
La Jolla, CA 92037
858-646-3100 ext.3620

Claudia Hayward-Costa wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Dear Experts,
}
} I am working with leukemia cells in suspension and spin them
} down in conical BEEM capsules prior to fixation. All subsequent
} steps (osmication, dehydration, infiltration with epon 812) are
} performed in these tubes.
} Unfortunately the blocks are very difficult to cut. The tips are with
} holes and inside the blocks there are "craters" and spongy areas
} that make it very difficult to cut proper sections.
} The pellets are minute, therefore I like the convenience of the
} conical tubes but I have to cut "big" chunks away with the razor
} blade before I get reasonable sections which makes it a quite
} wasteful job.
} The absolute EtOH is dried over Cu2SO4 in a dialysis tube and the
} cells are in filtrated twice within 24 hours.
}
} I know that one way out of this dilemma would be to handle the
} pellet as a tissue - but the experiment is so time consuming that I
} still do not want to risk having a crumbling pellet after 2 days work.
}
} Thanks in advance for your kind help.
}
} Claudia
}
} Dr. C. Hayward-Costa
} School of Life Sciences
} Kingston University
} Penrhyn Road, Kingston upon Thames
} Surrey KT1 2EE, UK
} ++44(0)181 547 2000 x 2240
} Email: c.hayward-at-kingston.ac.uk

--








From: Shane Collins :      kshanec-at-gte.net
Date: Mon, 13 Dec 1999 18:13:21 -0800
Subject: barcoded microscope glass slides

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Does anyone know if there are available in bulk glass microscope slides with
a barcode pre printed?

Thanks in advance,
Shane Collins
Scientific Instruments






From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Mon, 13 Dec 1999 19:35:04 -0800
Subject: Re: B: BEEM capsulas,embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Claudia, hello

I am working with HELA and COS7 cells in suspension. I would prefer to fix
(GA and OsO4 postfixation) them in suspension in 1.5 ml Eppendorf tubes.
Usually I collect cells by centrifugation of the suspension in the standard
cell culture 15 ml tubes at the low speed (not more than 1000 rpm on the
GLC-2 backet rotor centrifuge, 10-15 min) and resuspend them in the 0.5-1
ml fixer solution. Than I transferred cells suspension into the 1.5 ml
Eppendorf tubes and performs all manipulations in those tubes (with some
centrifugation between steps to hold precipitate in place if necessary).
At such conditions cells do not form a solid precipitate, therefore they
are more accessible to the solvents. If it necessary at the last step, I
transfer cells into BEEM capsules with pure plastic media, centrifuge it at
higher than before speed (say 2000 rpm, 20-30 min) and than polymerize it.
I find, BEEM capsules is less suitable for cells processing mostly because
a small volume. For approximately 50 um precipitate volume I prefer to use
about 500-700 um of the each solvent. It may be important at dehydrating
steps when in the small volume it is most likely to "re-hydrate" the
solvent (and sample as well) from air's moisture.

Good luck.

Sergey

} Dear Experts,
}
} I am working with leukemia cells in suspension and spin them
} down in conical BEEM capsules prior to fixation. All subsequent
} steps (osmication, dehydration, infiltration with epon 812) are
} performed in these tubes.
} Unfortunately the blocks are very difficult to cut. The tips are with
} holes and inside the blocks there are "craters" and spongy areas
} that make it very difficult to cut proper sections.
} The pellets are minute, therefore I like the convenience of the
} conical tubes but I have to cut "big" chunks away with the razor
} blade before I get reasonable sections which makes it a quite
} wasteful job.
} The absolute EtOH is dried over Cu2SO4 in a dialysis tube and the
} cells are in filtrated twice within 24 hours.
}
} I know that one way out of this dilemma would be to handle the
} pellet as a tissue - but the experiment is so time consuming that I
} still do not want to risk having a crumbling pellet after 2 days work.
}
} Thanks in advance for your kind help.
}
} Claudia
}
} Dr. C. Hayward-Costa
} School of Life Sciences
} Kingston University
} Penrhyn Road, Kingston upon Thames
} Surrey KT1 2EE, UK
} ++44(0)181 547 2000 x 2240
} Email: c.hayward-at-kingston.ac.uk

--





_________________________________

Sergey Ryazantsev Ph. D.
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu
http://www.bol.ucla.edu/~sryazant







From: ctschristopher :      ctschristopher-at-samiot.uct.ac.za
Date: Tue, 14 Dec 1999 10:14:26 +0300
Subject: B: Re BEEM Capsules embedding problems.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Claudia

It sounds like you have a problem with residue of fluids in your beem
capsules which could be preventing decent impregnation of your pellets. You
do not mention the source of your cells, but a few years ago when I worked
at Red Cross Children's Hospital in Cape Town, we had easy techniques for
both peripheral blood samples and aspirated cells.

Peripheral blood was centrifuged in capillary tubes to separate the cells
(this was done by our haem lab before they sent them to us). A score was
made on the glass of the cap tubes just below the buffy coat and then by
placing the broken end into a vial of fixative (we used cacodylate buffered
1.5% gluteraldehyde) and tapping gently on the other end, the plug of cells
were released. They were fixed for at least 2 hours and then processed
(whole, if 2mm long or less) in an automatic tissue processor (Shandon Lynx)
with our routine specimens. I would recommend that if you use an auto
processor you should line the basket to stop the specimens from being washed
out. We used to use a small piece of lens cleaning tissue with all of our
small specimens. These blocks are easy to cut although you sometimes had to
trim a bit to get through the layer of platelets as it is nearly impossible
to orientate the plugs.

For aspirated cells, we received cells in 1.5% glute which we centrifuged in
eppendorf tubes and removed all of the supernatant fluid. We then put a drop
of 5% - 10% gelatine or agar (just warmed enough to melt) and slightly
loosened the pellet slightly with an orange stick so that the cells were
concentrated at the base of the agar plug. This was then set in the fridge
and the pellets were trimmed and processed in the tissue processor with the
routine specimens. These blocks tended to be slightly soft but they still
sectioned perfectly.

Hope this helps.

Regards

Phil





From: rfelten-at-Macdermid.com
Date: Tue, 14 Dec 1999 08:04:32 -0500
Subject: What about a Annual "Printoff"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Rick Felten-at-MACDERMID
12/14/99 08:04 AM
Every few months printers are discussed on the listserver. I was wondering
if it wouldn't be a bad idea if we had an annual "printoff". We could have
two or three categories ( {$500 and } $500 or what ever). We could have a
standard image printed to a standard size. The submittor could record the
print time. The standard image could be a composite of features testing
resolution, nearly full dark regions, and nearly full white regions. Let
me know what you think.
Ric







From: jorge.duarte-at-amd.com
Date: Tue, 14 Dec 1999 08:16:50 -0600
Subject: Voltage Contrast

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi
I am trying to set for the first time a passive Voltage Contrast
analysis for contact 1, metal 1 and metal 2 using a FIB and I want to know
how is it done?, beside the FIB do I need any other equipment's? and is
there any publications that I can search about the analysis itself? Thank
you for your help.

AMD { {...} }

Jorge Duarte
Materials Engineer
F25 PCAL
Advanced Micro Devices
Jorge.duarte-at-amd.com
(512) 602-1431
1-800-538-8450 x51431







From: Paula S. Scott STP 727/896-8626 :      Paula.S.Scott-at-dep.state.fl.us
Date: Tue, 14 Dec 1999 09:29:22 -0500 (EST)
Subject: Re: B: BEEM capsulas,embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I usually have no problem with fixation/embedding of cell pellets. However
sometimes material is lost during the various steps. I find that if you embed
the pellet in agar after fixation that the results are often improved and
material is not lost. This will also eliminate the air problems you discribe.
I hope that this helps and good luck.






Paula Scott
Fish and Wildlife Conservation Commission
Florida Marine Research Institute
100 8th Avenue SE
St. Petersburg, FL 33701
(727) 896-8626
Fax (727) 823-0166






From: Stephen McCartney :      stmccart-at-vt.edu
Date: Tue, 14 Dec 1999 09:28:28 -0500
Subject: repair of vacuum pump

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good Morning: I have an Alcatel vacuum pump in need of repair. Can anyone
recommend somewhere in/near VA I can use. thanks.


------------------------------
Stephen McCartney
Research Associate
Virginia Tech
Materials Institute
2108 Hahn Hall
Blacksburg, VA 24061-0344
USA

TEL: 540-231-9765
FAX: 540-231-8517
------------------------------





From: Rujida Leepipattanawit :      leepipat-at-pilot.msu.edu
Date: Tue, 14 Dec 1999 10:23:33 -0500 (EST)
Subject: PET/PEN blend inquiry

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,


I am working on the PET/PEN blends and trying to get
the microstructure images of this Polymer blends by TEM.
However, By Ruthenium staining (Montezinos et al 1985)
of ultrathin sections,no phase differences were visible.
I espected to obtain the phase seperate of PET phase (domain phase)
and PEN phase.
I also tried SEM and found no seperation.

Please give any suggestion about "the sample Preparation".

regards,

Rujida Leepipattanawit







From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Tue, 14 Dec 1999 09:20:38 -0600
Subject: Re: What about a Annual "Printoff"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I agree with you that such a side-by-side comparison would be a good idea.
Indeed, a few years ago there was such a "Print-Off" that was displayed in
the computer lab at the MSA/MAS meeting. Nestor or someone had prepared a
test image with several sub-images to test several modes of imaging. I
think the image was available via the MSA web server. However, I don't
remember seeing one these last couple of years. Maybe it should be an
official and permanent part or the annual meeting.

Warren S.

At 08:04 AM 12/14/1999 -0500, you wrote:
} Rick Felten-at-MACDERMID
} 12/14/99 08:04 AM
} Every few months printers are discussed on the listserver. I was wondering
} if it wouldn't be a bad idea if we had an annual "printoff". We could have
} two or three categories ( {$500 and } $500 or what ever). We could have a
} standard image printed to a standard size. The submittor could record the
} print time. The standard image could be a composite of features testing
} resolution, nearly full dark regions, and nearly full white regions. Let
} me know what you think.
} Ric






From: Peter Steele :      STEELEP-at-allkids.org
Date: Tue, 14 Dec 1999 14:26:44 -0500
Subject: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


A recent posting reminded me of a problem that infrequently occurs within =
our lab. In fully processed specimens, the epon blocks will develop =
bubbles as they harden. At times there are so many bubbles that the =
specimen will be displaced from an oriented position. To our best =
efforts, we have not been able to identify what causes these bubbles. =
There does not seem to be a correlation with the age of any one compound =
or mold type. However, making up all new components will often (not =
always) get rid of the bubbles. We use vacuum, both after mixing epon =
components and again before the epon is cured (i.e., with the tissue =
embedded in epon). The epon is EMS Embed-812. Processing is on a Lynx, =
with glut, cacodylate buffers, osmium, graded alcohols, propylene oxide, =
and uncatalyzed epon. Typically, after processing, specimens sit under =
vacuum for 2-4 hours in epon with catalyst (BDMA) before being placed in =
the oven. Oven temperature is constant (70C), and curing is overnight. I =
have always been curious as to what causes these bubbles and how to avoid =
their formation.

TIA

P. Steele, Ph.D.
Pathology & Lab. Medicine
All Children's Hospital






From: Alan W. Nicholls :      nicholls-at-uic.edu
Date: Tue, 14 Dec 1999 14:42:06 -0600
Subject: Heating stages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have recently received a Gatan double tilt heating stage capable of
reaching 1000 degC. On our first trial run in the JEOL JEM-3010 a stage
mechanical limit alarm started above 550 deg C despite the stage being near
to the centre. The stage was free to move in any direction so we do not
think it was a real alarm and it slowed down and stopped as the temperature
was reduced below 500degC. I was wondering if anyone else out there has
had a similar problem?

Secondly someone wants to look at a cross sectional specimen above 500 deg
C. Does anyone know of a glue/cement that can be used safely upto 1000 deg C!

Regards

Alan W Nicholls, PhD
Manager - Electron Microscopy Service
Research Resources Center - East (M/C 337)
Room 100 Science and Engineering South Building
The University of Illinois at Chicago
845 West Taylor St
Chicago, IL 60607-7058

Tel: 312 996 1227
Fax: 312 996 8091
Office: Room 110

Web site www.rrc.uic.edu





From: Nestor J. Zaluzec :      zaluzec-at-aaem.amc.anl.gov
Date: Tue, 14 Dec 1999 15:11:37 -0600
Subject: Heating Stage Temps...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Alan

I have used a Gatan DT Heating stage (on a EM420T) to ~ 800 C.
without problems. The sample was not however a X-section but
a 3 mm plan-view sample.

Nestor
Your Friendly Neighborhood SysOp.

==================================================================
Nestor J. Zaluzec
Materials Science Division
Building 212
Argonne National Lab
9700 S. Cass Ave
Argonne, Illinois 60439 USA
Tel: 630-252-7901, Fax: 630-252-4798
Email: Zaluzec-at-aaem.amc.anl.gov
==================================================================
TPMLab: http://tpm.amc.anl.gov
MMSite: http://www.amc.anl.gov
==================================================================

The box said "This program requires Win 95/98/NT or better..." so I bought
a G3 Mac

==================================================================







From: Barbara Plowman :      Bplowman-at-sfmail.dental.uop.edu
Date: Tue, 14 Dec 1999 13:19:04 -0800
Subject: Bio/BEEM capsules, embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Claudia,
I have processed cultured cells successfully by using polypropylene =
centrifuge tubes or microcentrifuge tubes (a little bigger than the BEEM =
capsules you have used). I spin after each step and infiltrate in either =
Spurr or LRWhite. My last steps are to transfer each infiltrated sample =
to the BEEM capsules. Then, I place the BEEM capsules in microcentrifuge =
tubes piggyback(you may have to cut the microcentrifuge tubes with cutters =
to make them fit) and spin them once more before placing them in the oven. =
It's a long prep, but I've had very good results.






From: Hendrik O. Colijn :      colijn.1-at-osu.edu
Date: Tue, 14 Dec 1999 16:24:00 -0500
Subject: Re: Heating stages

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Alan,

Contact alarms usually detect current flow between the sample rod and the
rest of the microscope. Could you be getting sufficient thermal electron
emission (perhaps from the heating wires) to trip the sensor?

Cheers,
Henk

At 02:42 PM 12/14/99 -0600, Alan W. Nicholls wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Hendrik O. Colijn colijn.1-at-osu.edu
Campus Electron Optics Facility Ohio State University
(614) 292-0674 http://web.ceof.ohio-state.edu
Murphy's Law: "If anything can go wrong, it will."
Commentary: "Murphy was an optimist."






From: micro-at-ldeo.columbia.edu (Dee Breger)
Date: Tue, 14 Dec 1999 16:27:48 -0500
Subject: Re: What about a Annual "Printoff"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I like this idea a lot

}
} Rick Felten-at-MACDERMID
} 12/14/99 08:04 AM
} Every few months printers are discussed on the listserver. I was wondering
} if it wouldn't be a bad idea if we had an annual "printoff". We could have
} two or three categories ( {$500 and } $500 or what ever). We could have a
} standard image printed to a standard size. The submittor could record the
} print time. The standard image could be a composite of features testing
} resolution, nearly full dark regions, and nearly full white regions. Let
} me know what you think.
} Ric

Dee Breger
Mgr. SEM/EDX Facility
Lamont-Doherty Earth Observatory
61 Route 9W
Palisades, NY 10964 USA
T: 914/365-8640
F: 914/365-8155
E: micro-at-ldeo.columbia.edu

www.aspp.com/gallery/index_archive4.html
www.discovery.com/area/science/micro/micro1.html







From: schubelk-at-mail.lafayette.edu (Kathryn Schubel)
Date: Tue, 14 Dec 1999 16:44:30 -0400
Subject: Ion mill

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for a used ion mill for preparation of TEM samples. If anyone
has an ion mill that they are looking to get rid or knows of anyone who is
please contact me.

Thanks.

Kathryn Schubel - Assistant Professor of Geology
Department of Geology and
Environmental Geosciences
Lafayette College
Easton, PA 18042
U.S.A.

(610) 330-5194 voice
(610) 330-5193 dept. office
(610) 330-5717 FAX
schubelk-at-lafayette.edu







From: DrJohnRuss-at-aol.com
Date: Tue, 14 Dec 1999 16:43:11 EST
Subject: multi-tiff image format

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


This is directed in particular at any one using an Olympus or Bio-Rad
confocal microscope, which I am told save their images in a multi-image TIFF
format. If you would contact me directly by e-mail I would appreciate it. I
am trying to get my hands on a couple of representative images, contents not
important, to see which of the many possible tiff file variants are in use.
If anyone else is using multi-tiffs from some other brand of machine, I'd
like to know that as well.

Thanks
John Russ
John_Russ-at-NCSU.edu or DrJohnRuss-at-AOL.com





From: Jaci Lett :      jmlett-at-cid.wustl.edu
Date: Tue, 14 Dec 1999 16:05:48 -0600
Subject: LM: HM 505 cryostat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We have an HM 505 cryostat that may need parts or need to be sent in for
repairs. I recall seeing something recently about Microm-Heidelberg
equipment being handled by Richard Allan Scientific. Is this true for all
HM equipment? We purchased this cryostat through a Zeiss vendor (who no
longer handles Zeiss) and we'd really like to make sure that we deal with
the proper company.

Who is actually handling repairs and parts for HM cryostats? I'll be
contacting Zeiss soon, but in the meantime, I'd like some warning as to what
I'll run into.

Please contact me personally,

Jaclynn M. Lett jmlett-at-cid.wustl.edu

Fay and Carl Simon Center for the Biology of Hearing and Deafness
Central Institute for the Deaf
818 S. Euclid Ave.
St. Louis, MO 63110

voice: 314-977-0257 fax: 314-977-0030






From: Walck. Scott D. :      walck-at-ppg.com
Date: Tue, 14 Dec 1999 18:25:07 -0500
Subject: RE: Heating stages -The glue

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Alan,
Check out the MRS TEM prep book I, vol 115, p.126, (1988). Eric Fiore and
Rodney Herring have a paper in there that describes a way of going up to
1300C for cross sectional samples. The "glue" that they use is Ceramabond
569 from Aremco Products, Inc.

-Scott


Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)


"The opinions expressed are those of Scott D. Walck and not of PPG
Industries, Inc. nor of any PPG-associated companies."
--


} -----Original Message-----
} From: Alan W. Nicholls [mailto:nicholls-at-uic.edu]
} Sent: Tuesday, December 14, 1999 3:42 PM
} To: microscopy-at-Sparc5.Microscopy.Com
} Subject: Heating stages
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.} html
}
}
} --------------------------------------------------------------
} ---------.
}
}
} We have recently received a Gatan double tilt heating stage capable of
} reaching 1000 degC. On our first trial run in the JEOL
} JEM-3010 a stage
} mechanical limit alarm started above 550 deg C despite the
} stage being near
} to the centre. The stage was free to move in any direction
} so we do not
} think it was a real alarm and it slowed down and stopped as
} the temperature
} was reduced below 500degC. I was wondering if anyone else
} out there has
} had a similar problem?
}
} Secondly someone wants to look at a cross sectional specimen
} above 500 deg
} C. Does anyone know of a glue/cement that can be used safely
} upto 1000 deg C!
}
} Regards
}
} Alan W Nicholls, PhD
} Manager - Electron Microscopy Service
} Research Resources Center - East (M/C 337)
} Room 100 Science and Engineering South Building
} The University of Illinois at Chicago
} 845 West Taylor St
} Chicago, IL 60607-7058
}
} Tel: 312 996 1227
} Fax: 312 996 8091
} Office: Room 110
}
} Web site www.rrc.uic.edu
}





From: PESTOEM-at-aol.com
Date: Tue, 14 Dec 1999 18:42:14 EST
Subject: Jeol 1200 EX

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To all Microscopists, does anyone have a complete good alignment procedure
for the Jeol 1200. We would very much appreciate a copy.
Thank you very much. Peter Stolzenberg, Pesto Inc. 215-699-6160
Fax 215-699-5275





From: Maria Lucia Ribeiro Caldas :      caldasml-at-amcham.com.br
Date: Tue, 14 Dec 1999 22:11:05 -0200
Subject: Second chance!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all

This is a second chance. I surely need the formula for chromotrope
aniline blue and syrius red. It's very important to me at the moment,
so I will be able to
get some improvement in my transplant kidney bx diagnosis. It's
Christmas time, a time for giving, please help me!

Don't need to be stressed I will post to Pathol....

Thanks
Lucia Caldas
Rua Pres. Backer 234/604 BII
Icarai Niteroi RJ
Brasil 24220-041






From: Gary Liechty :      garyliechty-at-att.net
Date: Tue, 14 Dec 1999 16:50:18 -0800
Subject: Re: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Mr. Steele,

Moisture is often caught within the sample, either in an air pocket or as=
it exists in the material. During the curing/heating of the Epoxy, the =
moisture vaporizes into a gas, forming a bubble. These bubbles do not es=
cape the mount because the Epoxy is too thick at that point.

Thoroughly drying the sample is the only way to ensure this does not happ=
en.

I hope this helps.

Sincerely,

Gary Liechty


Peter Steele wrote:

} -----------------------------------------------------------------------=
-
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=

} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Co=
m
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.htm=
l
} -----------------------------------------------------------------------=
=2E
}
} A recent posting reminded me of a problem that infrequently occurs with=
in our lab. In fully processed specimens, the epon blocks will develop b=
ubbles as they harden. At times there are so many bubbles that the speci=
men will be displaced from an oriented position. To our best efforts, we=
have not been able to identify what causes these bubbles. There does no=
t seem to be a correlation with the age of any one compound or mold type.=
However, making up all new components will often (not always) get rid o=
f the bubbles. We use vacuum, both after mixing epon components and agai=
n before the epon is cured (i.e., with the tissue embedded in epon). The=
epon is EMS Embed-812. Processing is on a Lynx, with glut, cacodylate b=
uffers, osmium, graded alcohols, propylene oxide, and uncatalyzed epon. =
Typically, after processing, specimens sit under vacuum for 2-4 hours in =
epon with catalyst (BDMA) before being placed in the oven. Oven temperat=
ure is constant (70C), and curing is overnight. I
} have always been curious as to what causes these bubbles and how to avo=
id their formation.
}
} TIA
}
} P. Steele, Ph.D.
} Pathology & Lab. Medicine
} All Children's Hospital

--
Gary Liechty
Product Application Specialist

Allied High Tech Products, Inc.
2376 E. Pacifica Place
Rancho Dominguez, CA 90220

310-635-2466
800-675-1118 US Only
310-762-6808 Fax

www.alliedhightech.com

Products and Equipment for Metallurgical Sample Preparation







From: Gary Liechty :      garyliechty-at-att.net
Date: Tue, 14 Dec 1999 16:51:42 -0800
Subject: Re: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Mr. Steele,

Moisture often exists within a sample, in an air pocket or as a component=
of the sample/material. During the curing/heating of the Epoxy, the moi=
sture vaporizes into a gas, forming a bubble. These bubbles do not escap=
e the mount because the Epoxy is too thick at that point.

Thoroughly drying the sample is the only way to reduce this effect.

I hope this helps.

Sincerely,

Gary Liechty


Peter Steele wrote:

} -----------------------------------------------------------------------=
-
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=

} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Co=
m
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.htm=
l
} -----------------------------------------------------------------------=
=2E
}
} A recent posting reminded me of a problem that infrequently occurs with=
in our lab. In fully processed specimens, the epon blocks will develop b=
ubbles as they harden. At times there are so many bubbles that the speci=
men will be displaced from an oriented position. To our best efforts, we=
have not been able to identify what causes these bubbles. There does no=
t seem to be a correlation with the age of any one compound or mold type.=
However, making up all new components will often (not always) get rid o=
f the bubbles. We use vacuum, both after mixing epon components and agai=
n before the epon is cured (i.e., with the tissue embedded in epon). The=
epon is EMS Embed-812. Processing is on a Lynx, with glut, cacodylate b=
uffers, osmium, graded alcohols, propylene oxide, and uncatalyzed epon. =
Typically, after processing, specimens sit under vacuum for 2-4 hours in =
epon with catalyst (BDMA) before being placed in the oven. Oven temperat=
ure is constant (70C), and curing is overnight. I
} have always been curious as to what causes these bubbles and how to avo=
id their formation.
}
} TIA
}
} P. Steele, Ph.D.
} Pathology & Lab. Medicine
} All Children's Hospital

--
Gary Liechty
Product Application Specialist

Allied High Tech Products, Inc.
2376 E. Pacifica Place
Rancho Dominguez, CA 90220

310-635-2466
800-675-1118 US Only
310-762-6808 Fax

www.alliedhightech.com

Products and Equipment for Metallurgical Sample Preparation







From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Tue, 14 Dec 1999 15:10:39 -1000 (HST)
Subject: Saline for Panaeus and Macrobrachium

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My collaborators would like to know the blood osmolarity of Panaeus
vanamei and of Macrobrachium. They are interested in finding good
salines for Panaeus and Macrobrachium that will make their dissected parts
happy for physiological experiments, and also a favorite buffer for
fixation of these happy parts for electron microscopy. Long ago I used to
use Dalton's saline for physilogical experiments on crab parts, but I'm
wondering if any of you have other opinions. For fixation I generally use
4% glut in 0.1M sodium cacodylate with 0.35M sucrose for the marine stuff,
but I'm willing to entertain other recipes as well.

Thanks in advance for any suggestions!

Mele Kalikimaka,
Tina

http://www.pbrc.hawaii.edu/bemf/microangela

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************






From: Marek Malecki :      mmm-at-biomail.ucsd.edu
Date: Tue, 14 Dec 1999 20:21:35 -0600
Subject: Ultramicrotomist wanted - life sciences.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At the Molecular Imaging Laboratories at the University of California , San
Diego, there is an immediate opening for an ultramicrotomist. A candidate
should be definitely hands-on familiar with Reichert and Sorvall microtomes
as well as with Epon, Spurr, and Araldite embedments. Experience with
sectioning of frozen sucrose infused samples is a big plus, but it can be a
subject of training on the site.
This is an equal employment opportunity. Minorities are strongly encouraged
to apply. Starting salary ~$24000.- Candidates are welcome to contact me
with any questions.

Marek Malecki, M.D., Ph.D.
Director and Principal Investigator

Molecular Imaging Laboratories
University of California at San Diego

address: 1500 Suite Bonner Hall, 9500 Gilman Drive, La Jolla, CA 92093-0368
phone: 8588223373
fax: 8588223715
pager: 8586161420
e.mail: mmm-at-ucsd.edu
e.pager: 1620024619-at-alphapage.airtouch.com
www site: http://mil.ucsd.edu/
ftp site: mil.ucsd.edu







From: Brian Gortney :      gortn-at-earthlink.net
Date: Tue, 14 Dec 1999 21:27:04 -0500
Subject: LM,Aus-Jena Manuals

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings:
I recently purchased the Aus-Jena Laboval scope that I Asked your
opinion of
last month.I am overjoyed with it.It is nearly new and looks and works as if
it has been
hardly used.I have a few questions some of you may be able to help me with ?
The unit
is approx.20 years old:
1.) Does anyone have any idea where I can find a factory manual on use &
maintenance
in English? or a reprint ?

2.) The unit has an accessory polarizer unit between the trinocular eyepiece
and the turrent,
it is marked 1.25X. I presume this means it increases effective
magnification by 1.25.Is this
correct ? The Trinocular head has a camera mount and its marked 1X at the
camera fitting,
and 1.6X at the eyepiece binocular unit.Would I multiply the objective power
by both 1.25 and
1.6 to figure the total working magnification ?

3.) Any Idea where I can get an English manual on the polarizer.I have a
wooden fitted box
full of many different lens plates & sliders etc. but no instructions or
anything.Soon I will figure
it out however,I would really like to have a manual for this accessory.

4.) I also have a separate darkfield condenser unit,any idea on manuals or
instructions ?

Thank you all for taking the time to read this,and special thanks to all who
can help point me in
the proper direction.
Brian Gortney

gortn-at-earthlink.net

PS: I purchased the unit from Mcbain Instruments in Chatsworth Ca. They are
a Leica dealer
and service facility.I am well pleased with them. I would not hesitate to
recommend them.
Gerry Burke has been most helpful and knowledgeable.(I have no affiliation
with Mcbain)






From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Tue, 14 Dec 1999 17:16:07 -1000 (HST)
Subject: Re: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I find that I can dramatically reduce the chance of bubbles if I pre-dry
the empty capsules and their labels in the oven all day or overnight
before adding the samples and resin. I got onto this many years ago when
I realized I was getting bubbles off the paper labels as well, so I
decided they held moisture and/or gas. Works for me!

Good luck,
Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************






From: phil.swab-at-depsci.com (Phil Swab)
Date: Tue, 14 Dec 1999 19:50:05 -0800
Subject: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Peter et al,

I have a materials background, so I'm sure I do not understand the process
of embedding biological samples. That said, the bubbles could be a result
of outgassing your specimen directly in the epoxy. Also, outgassing
specimens and epoxy for 2 - 4 hours sounds excessive and could drive off
much of the typically aromatic catalyst used for free-radical addition
polymerization of epoxies. I typically do not outgas my epoxy (Spurrs), I
try to mix it very gently so as not to mix air into it.

On occasion, for very porous problem materials, I will put the epoxy and
specimen under vacuum for at most 5 minutes. However, I do not place the
specimen in the epoxy during this time, as the surface tension of liquids
will impede the evacuation process. I place the specimen on a piece of
foil attached to the lip of a small beaker of epoxy. After outgassing in a
small bell jar for a few minutes, I gently tap the base plate of the bell
jar to knock the specimen into the epoxy, and then leak the chamber back up
to atmospheric pressure. In this process you more efficiently evacuate the
interstitial spaces of the samples and use atmospheric pressure at 14 psi
to force epoxy into the evacuated voids. This has worked well for fine
sub-micron sized powder, fabric, sponge, paper, and catalysts.

When embedding pre-processed wet tissue, it is important to first outgas
the sample in the lowest molecular weight liquid. Infiltration through
serial dilutions of solvent to viscous epoxy then occurs primarily by
diffusion, minimizing tissue trauma and damage.

The vacuum pulled should not reach pressures so low that active boiling of
the epoxy occurs - a gentle release of small bubbles is all the is needed,
and then only to indicate outgassing. There's no reason to outgas
dissolved gases - leave them dissolved, it's only necessary to evacuate the
voids and trapped gas in your sample prior to embedding.

Cheers,

Phil Swab
Engineering Development
Deposition Sciences Inc.
Santa Rosa, CA
707-566-3718
phil.swab-at-depsci.com

-----Original Message-----
} From: Peter Steele [SMTP:STEELEP-at-allkids.org]
Sent: Tuesday, December 14, 1999 11:27 AM
To: Microscopy-at-Sparc5.Microscopy.Com


A recent posting reminded me of a problem that infrequently occurs within
our lab. In fully processed specimens, the epon blocks will develop
bubbles as they harden. At times there are so many bubbles that the
specimen will be displaced from an oriented position. To our best efforts,
we have not been able to identify what causes these bubbles. There does
not seem to be a correlation with the age of any one compound or mold type.
However, making up all new components will often (not always) get rid of
the bubbles. We use vacuum, both after mixing epon components and again
before the epon is cured (i.e., with the tissue embedded in epon). The
epon is EMS Embed-812. Processing is on a Lynx, with glut, cacodylate
buffers, osmium, graded alcohols, propylene oxide, and uncatalyzed epon.
Typically, after processing, specimens sit under vacuum for 2-4 hours in
epon with catalyst (BDMA) before being placed in the oven. Oven
temperature is constant (70C), and curing is overnight. I have always been
curious as to what causes these bubbles and how to avoid their formation.

TIA

P. Steele, Ph.D.
Pathology & Lab. Medicine
All Children's Hospital








From: r.cross-at-ru.ac.za
Date: Wed, 15 Dec 1999 09:19:31 +0200
Subject: Re: What about a Annual "Printoff"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello all

} I agree with you that such a side-by-side comparison would be a good
} idea.

Yes, this is a very good idea.

Another interesting variable is the paper on which the images are
printed. At the Microscopy Society of Southern Africa's recent
meeting there was a very informative display of images printed
(mainly by different HP and Epson inkjets, if I remember correctly)
on a wide variety of papers. This also included SEM images of the
paper. I don't remember which turned out to be the "best" paper (or
"best value for money" paper) but there certainly was a wide
difference between the quality of the images and the fine structure
of the papers that were examined. Something that was
immediately obvious to me was the wide variety of contrast
produced by different papers.

Warm regards from us all down here (in the high 30's C this week)!







Robin H Cross
Director : EM Unit, Rhodes University, Grahamstown, South Africa
R.Cross-at-ru.ac.za
tel: +27 46 603 8168 - fax: +27 46 622 4377
http://www.ru.ac.za/emu/index.htm

** remember that ICEM-15 in 2002 is in Durban, South Africa **





From: Chris Gilpin :      cgilpin-at-man.ac.uk
Date: Wed, 15 Dec 1999 11:55:40 -0000
Subject: LINK AN10000 replacement hard disk drive - source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi
I have an AN10000 EDX system. The hard disk has gone down. I usually use NEC
D3142 20Mb replacements. Sadly my supplier can no longer source this drive.
Does anyone have a suggestion for an alternative drive.

Many thanks

Chris


Chris Gilpin
Experimental Officer
Biological Sciences EM Unit
G452 Stopford Building
Oxford Road
Manchester
M13 9PT
phone +44 0161 275 5170
Fax +44 0161 275 5171
http://www.empgu.man.ac.uk






From: Malcolm Haswell :      malcolm.haswell-at-sunderland.ac.uk
Date: Wed, 15 Dec 1999 11:08:22 +0000
Subject: Re: B: BEEM capsulas,embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Claudia

have you considered the possibility that you are simply spinning down
any hard debris to the bottom of your capsule which can cause holes in
sections and general problems with embedding. This seems to be a classic
problem with small pellets especially where you have to spin after each
solution and the debris may come from a variety of sources:
the sample
crystals or deposits in solution (maybe your Cu2SO4 is leaking out of
the dialysis tube)
bits of broken glass from pipettes etc

The simplest answer is just to trim past this layer, but you could try
filtering everything before use (0.2um filters on syringes) and avoiding
the use of glassware with these difficult pellets (disposable plastic
pasteurs etc). Otherwise encapsulate in agar or try pelleting in
haematocrit glass capillaries as others have suggested.

Oh and merry Christmas and happy new Millennium to one and all (the
above reply was just an excuse for this greeting really).

Malcolm

Malcolm Haswell
Electron Microscopy
School of Sciences
Fleming Building
University of Sunderland
SUNDERLAND SR1 3SD
Tyne and Wear
UK

Tel (0191) 515 2872
e-mail: malcolm.haswell-at-sunderland.ac.uk
-----------------------------------------------------
Claudia Hayward-Costa wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Dear Experts,
}
} I am working with leukemia cells in suspension and spin them
} down in conical BEEM capsules prior to fixation. All subsequent
} steps (osmication, dehydration, infiltration with epon 812) are
} performed in these tubes.
} Unfortunately the blocks are very difficult to cut. The tips are with
} holes and inside the blocks there are "craters" and spongy areas
} that make it very difficult to cut proper sections.
} The pellets are minute, therefore I like the convenience of the
} conical tubes but I have to cut "big" chunks away with the razor
} blade before I get reasonable sections which makes it a quite
} wasteful job.
} The absolute EtOH is dried over Cu2SO4 in a dialysis tube and the
} cells are in filtrated twice within 24 hours.
}
} I know that one way out of this dilemma would be to handle the
} pellet as a tissue - but the experiment is so time consuming that I
} still do not want to risk having a crumbling pellet after 2 days work.
}
} Thanks in advance for your kind help.
}
} Claudia
}
} Dr. C. Hayward-Costa
} School of Life Sciences
} Kingston University
} Penrhyn Road, Kingston upon Thames
} Surrey KT1 2EE, UK
} ++44(0)181 547 2000 x 2240
} Email: c.hayward-at-kingston.ac.uk





From: DayDawning-at-aol.com
Date: Wed, 15 Dec 1999 07:47:53 EST
Subject: Re: LM: HM 505 cryostat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi Jaclyn,
Richard-Allan has purchased Microm. Richard-Allan/Microm will be servicing
the equipment in the near future, but for service now please contact Zeiss
Service.

Dawn M. Truscott, HT(ASCP)
RAS/Microm Instrumentation Team





From: Keith Ryan :      KPR-at-wpo.nerc.ac.uk
Date: Wed, 15 Dec 1999 12:55:09 +0000
Subject: LM stereo optics problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi All

I am trying to help a student who has been bought a 10-40x stereo
zoom microscope with some image recording/analysis equipment. The
microscope is totally unbranded - nothing on the body or the eyepieces
(except for magnification).

The problem is an apparent mismatch of the lamp intensity in the
eyepieces. In the right ocular, the illumination is brighter
centrally, and it zooms up concentrically - that's fine. In the left
ocular, the brighter central zone zooms up to go out of the field of
view, to the left. A centred specimen zooms up concentrically in both
oculars. Any suggestions? Is it a phenomenon of the different height
of the bulb, which is directly below the stage, and the specimen? The
problem is apparently alleviated by putting a piece of velin tissue
(thin paper) under the specimen. But this introduces a patterned
background.

Keith Ryan
Marine Biological Association
Plymouth UK





From: Dr Klaus D. Jandt :      K.Jandt-at-bristol.ac.uk
Date: Wed, 15 Dec 1999 13:42:26 +0000 (GMT Standard Time)
Subject: Biomaterials SPM 2000 Call for papers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


1st CALL FOR PAPERS AND POSTERS

and

Invitation for Delegates

2nd International Conference on Scanning Probe Microscopy in
Biomaterials Science

23 June 2000

Holiday Inn Crowne Plaza Hotel
Bristol, England

Offical website: http://www.dent.bris.ac.uk/biomaterials/spm2000/

Although established as a tool in materials science and physics, scanning=
=20
probe microscopy (SPM) is at the beginning of its application in biomateria=
ls=20
science. On 2 April 1998 the first workshop entitled "Scanning Probe Micros=
copy=20
in Biomaterials Science, Dentistry and Medicine" was held in Bristol, UK.=
=20
What was planned to be a small workshop evolved to be an international conf=
erence with high calibre=20
delegates and speakers from all over the world. Encouraged by the success o=
f the meeting and supported
by international academics and industrial researchers we are organising a 2=
nd
conference.
=20
Since this first conference in 1998 more researchers have applied atomic fo=
rce=20
microscopy and related SPM methods in biomaterials science.=20
Therefore a definitive need for a broad scientific exchange between researc=
hers involved in these studies exists.
This is the purpose of The 2nd International Conference on Scanning Probe =
Microscopy in Biomaterials Science,=20
which will be hosted by the University of Bristol and Veeco Instruments Lim=
ited.

Contributions should cover, but are not limited to, the following areas:

Imaging of biomaterials surfaces (polymers, metals ceramics etc.)=20
Interfaces between biomaterials and biological materials (e.g. protein-biom=
aterial interfaces)=20
Investigation of local properties of biomaterials (mechanical, chemical etc=
.)=20
Structural change of biomaterials=20
Aspects of medicine and dentistry relevant for SPM (e.g. SPM on mineralised=
tissues or DNA)=20
Instrumental developments in SPM and combination with other methods in the=
=20
investigations of biomaterials

Deadlines and dates

1 September 1999: early registration starts
1 January 2000: registration starts
1 April 2000 deadline for abstract submission
1 June 2000 registration closes =96 late registration (at an increased fee =
rate) possible until the date of the conference.

Speakers (confirmed):=20
Saul Tendler, Nottingham, UK
Roger Marchant, Cleveland, OH, USA
Grayson W. Marshall, San Francisco, USA
Buddy D Ratner, Seattle, USA
Klaus Jandt, Bristol, UK
etc.

Poster and presentations sessions: delegates will be able to present poster=
s or=20
give 15 min. oral presentations.

Registration : http://www.dent.bris.ac.uk/biomaterials/spm2000/

I am looking forward to welcome you in Bristol

Yours sincerely

Klaus D Jandt

---------------------------------------------------------------------------=
-----
Dr. rer. nat. Klaus D. Jandt
Senior Lecturer in Dental Materials Science and Biomaterials
University of Bristol, Department of Oral and Dental Science
Lower Maudlin Street, Bristol, BS1 2LY, UK
Phone: ++ 44 (0) 117 9 28 44 18, Fax: ++ 44 (0) 117 9 28 47 80
Internet: K.Jandt-at-bris.ac.uk
WWW: http://www.dent.bris.ac.uk/Biomaterials/kdj.htm
"We make Biomaterials Science work!"




















From: Malcolm Haswell :      malcolm.haswell-at-sunderland.ac.uk
Date: Wed, 15 Dec 1999 13:46:28 +0000
Subject: Re: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have certainly experienced the problems that Phil mentions. In fact I
would only pump down a chamber slowly and bleed air back in if excessive
bubbling occurred. Most specimens only needed a few minutes of
outgassing with Spurr's.
Perhaps part of your problem is due to the higher viscosity of Epon
which as it warms becomes less viscous and releases trapped bubbles.

Malcolm

Malcolm Haswell
Electron Microscopy
School of Sciences
Fleming Building
University of Sunderland
SUNDERLAND SR1 3SD
Tyne and Wear
UK

Tel (0191) 515 2872
e-mail: malcolm.haswell-at-sunderland.ac.uk


Phil Swab wrote:
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} Peter et al,
}
} I have a materials background, so I'm sure I do not understand the process
} of embedding biological samples. That said, the bubbles could be a result
} of outgassing your specimen directly in the epoxy. Also, outgassing
} specimens and epoxy for 2 - 4 hours sounds excessive and could drive off
} much of the typically aromatic catalyst used for free-radical addition
} polymerization of epoxies. I typically do not outgas my epoxy (Spurrs), I
} try to mix it very gently so as not to mix air into it.
}
} On occasion, for very porous problem materials, I will put the epoxy and
} specimen under vacuum for at most 5 minutes. However, I do not place the
} specimen in the epoxy during this time, as the surface tension of liquids
} will impede the evacuation process. I place the specimen on a piece of
} foil attached to the lip of a small beaker of epoxy. After outgassing in a
} small bell jar for a few minutes, I gently tap the base plate of the bell
} jar to knock the specimen into the epoxy, and then leak the chamber back up
} to atmospheric pressure. In this process you more efficiently evacuate the
} interstitial spaces of the samples and use atmospheric pressure at 14 psi
} to force epoxy into the evacuated voids. This has worked well for fine
} sub-micron sized powder, fabric, sponge, paper, and catalysts.
}
} When embedding pre-processed wet tissue, it is important to first outgas
} the sample in the lowest molecular weight liquid. Infiltration through
} serial dilutions of solvent to viscous epoxy then occurs primarily by
} diffusion, minimizing tissue trauma and damage.
}
} The vacuum pulled should not reach pressures so low that active boiling of
} the epoxy occurs - a gentle release of small bubbles is all the is needed,
} and then only to indicate outgassing. There's no reason to outgas
} dissolved gases - leave them dissolved, it's only necessary to evacuate the
} voids and trapped gas in your sample prior to embedding.
}
} Cheers,
}
} Phil Swab
} Engineering Development
} Deposition Sciences Inc.
} Santa Rosa, CA
} 707-566-3718
} phil.swab-at-depsci.com
}
} -----Original Message-----
} } From: Peter Steele [SMTP:STEELEP-at-allkids.org]
} Sent: Tuesday, December 14, 1999 11:27 AM
} To: Microscopy-at-Sparc5.Microscopy.Com
} Subject: Epon bubbles
}
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} A recent posting reminded me of a problem that infrequently occurs within
} our lab. In fully processed specimens, the epon blocks will develop
} bubbles as they harden. At times there are so many bubbles that the
} specimen will be displaced from an oriented position. To our best efforts,
} we have not been able to identify what causes these bubbles. There does
} not seem to be a correlation with the age of any one compound or mold type.
} However, making up all new components will often (not always) get rid of
} the bubbles. We use vacuum, both after mixing epon components and again
} before the epon is cured (i.e., with the tissue embedded in epon). The
} epon is EMS Embed-812. Processing is on a Lynx, with glut, cacodylate
} buffers, osmium, graded alcohols, propylene oxide, and uncatalyzed epon.
} Typically, after processing, specimens sit under vacuum for 2-4 hours in
} epon with catalyst (BDMA) before being placed in the oven. Oven
} temperature is constant (70C), and curing is overnight. I have always been
} curious as to what causes these bubbles and how to avoid their formation.
}
} TIA
}
} P. Steele, Ph.D.
} Pathology & Lab. Medicine
} All Children's Hospital





From: Pamela Neill :      Pamela.Neill-at-alconlabs.com
Date: Wed, 15 Dec 1999 07:55:53 -0600
Subject: Liver micrographs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


My niece's science project for fourth grade is liver and related diseases.
Does anyone have micrographs of normal or diseased liver that could be
shared with her class?
Please contact me off-list
Thank you very much
pamela.neill-at-alconlabs.com





From: Robert Santoianni :      Robert_Santoianni-at-emory.org
Date: Wed, 15 Dec 1999 08:40:40 -0500
Subject: "Epon" Bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Peter,
I also use EMbed-812, but polymerize overnight at 60 degrees. I do not
use any vacuum steps and I use DMP-30 instead of BDMA. I don't perform
any extra preparation steps to remove moisture from anything (and we're
in the SE where it's quite humid in the summer). I have never had a
problem with bubbles. I think that 70 degrees is excessive for
overnight polymerization. This may be the source of your problem. Call
Stacy Kirsch at Electron Microscopy Sciences to troubleshoot this
problem. If she or the chemistry people who make EMbed-812 can't
suggest a solution, I would be surprised! She's probably trying to
unclench her jaw right now, having seen your question!!!

Bob Santoianni
Emory University Hospital
Atlanta, GA





From: Wayne England :      wengland-at-ortech.on.ca
Date: Wed, 15 Dec 1999 09:08:20 -0500
Subject: equipment available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi all,

A colleague has asked for a recommendation for a camera to be used as
described below. I would appreciate any recommendations you can
make. Since this has been a topic before, perhaps it would be best to
reply directly to me.

Thanks so much for any help you can provide.

Damian Neuberger, Ph D
Research Scientist
Baxter Healthcare, Inc.
WG3-2S
P.O. Box 490
Round Lake, IL 60073
Tel: 847.270.5888
FAX: 847.270.5897

Microscope: Nikon Eclipse E-400 FITC epifluorescence with 10X and 60X
high-dry objectives.




In the midst of lab relocations we have decided to part with some equipment
that is not under heavy use.

Available is a Philips 400T STEM with EDAX capabilities which has functioned
well for us and an RMC cryo-ultramicrotome MT-6000 XL. The microtome is in
excellent condition (like new) with very little use over the years.

We have additional accessories and some parts available as well.

Please respond directly by e-mail or phone.

Prices are negotiable but must be reasonable.

Regards,

Wayne England


============================================
Wayne England
Manager, Physical Characterization
Bodycote ORTECH Inc.
2395 Speakman Drive, Mississauga, ON, L5K1B3
wengland-at-ortech.on.ca WEB: www.bodycote.com
905-822-4111 Ext.555 FAX:905-823-1446
============================================






From: Walck. Scott D. :      walck-at-ppg.com
Date: Wed, 15 Dec 1999 09:13:23 -0500
Subject: Re: Print-off idea ... **** or get off the pot.

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Nestor did this several years ago. He sent around a test image that was
fairly representative of the types of images found in microscopy labs.
Since he didn't jump into this thread, he probably doesn't want the hassle
of setting it up the comparison table again. What I suggest is that someone
that is interested in doing it and who is going to attend the M&M MM meeting
volunteer to take responsibility to do this. I'm sure that Nestor would
send the test image to that person for distribution to those who want to
participate. If you are the person who would like the responsibility,
contact Stuart McKernan (the M&M MM Program Chair) and ask to do it. I
suggest that you coordinate it with Nestor and John Mansfield. Stuart's
Email address is
mckernan-at-cems.umn.edu

If nobody want to do it, then let's drop the thread.

Just my two cents.

-Scott


Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)







From: Nestor J. Zaluzec :      zaluzec-at-Sparc5.Microscopy.Com
Date: Wed, 15 Dec 1999 08:15:17 -0600
Subject: Re: LINK AN10000 replacement hard disk drive - source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chris

I have replaced these drives with SSCSI HD drives from old MacIntosh computers
they work fine, and only have to be reformatted. Your hardware may also
permit } than 20Mb drives, different chip sets in the AN1000 may allow you to
use 40 / 80 Mb drives. You'll find out when you format them.

Also check your 5 V power supply. I've had alot of problems with the HD and
it turns out that the problem was sometimes the voltage levels. Just unplug
the floppy drive and stick a DVM in the power plug. You should get } 5 V.
if it drops below 5 then HD action will act as if the drive is dead.

A key thing to check is the connectors from the PS to the Bus. They are
silver coated and tarnish. Get a bit of metal polish and clean them off. On
my system this made a 0.25 V difference!

Nestor
Your Friendly Neighborhood SysOp.



} Hi
} I have an AN10000 EDX system. The hard disk has gone down. I usually use NEC
} D3142 20Mb replacements. Sadly my supplier can no longer source this drive.
} Does anyone have a suggestion for an alternative drive.
}
} Many thanks
}
} Chris
}
}
} Chris Gilpin
} Experimental Officer
} Biological Sciences EM Unit
} G452 Stopford Building
} Oxford Road
} Manchester
} M13 9PT
} phone +44 0161 275 5170
} Fax +44 0161 275 5171
} http://www.empgu.man.ac.uk








From: c j day :      wa5ekh-at-juno.com
Date: Wed, 15 Dec 1999 08:15:50 -0600
Subject: ESEM Installation and Survey(EMF and Mechanical)

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am trying to find someone to survey and install a ESEM + EDS in the
Dallas Area by the end of the year. We are going to evaluate the system,
so we are trying to minimize this installation cost. We will probably
move it again in about year. Any suggestions?
Jeff Day/JD
wa5ekh-at-juno.com







From: giacomo.parodi-at-it.abb.com ()
Date: Wed, 15 Dec 1999 08:20:18 -0600
Subject: LM: spirocheates:minimum magnification needed?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Email: giacomo.parodi-at-it.abb.com
Name: Giacomo

Question: Please tell me the minimum magnification needed
to recognise spirocheates without doubts.
I think of using darkfield and I have to
decide if I need to buy a standard condenser
plus a light stop or a specialized darkfield
condenser. I know that a light stop works
only at low magnification.

Thank you.


---------------------------------------------------------------------------







From: M.C. Guadalupe Nieto :      gnieto-at-tap-ecosur.edu.mx
Date: Wed, 15 Dec 1999 09:30:02
Subject: unsuscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please unsuscribe me,

Thanks.


MC. Ma. Guadalupe Nieto L=F3pez
Laboratorio de Microscop=EDa Electr=F3nica
ECOSUR Unidad Tapachula
Carr. Antiguo Aeropouerto Km 2.5
30700 Tapachula, Chiapas, Mexico.
Tel. (962) 81077, 81103
Fax. (962) 81015







From: Leona Cohen-Gould :      lcgould-at-mail.med.cornell.edu
Date: Wed, 15 Dec 1999 10:47:39 -0400
Subject: Re: Second chance!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all
}
} This is a second chance. I surely need the formula for chromotrope
} aniline blue and syrius red. It's very important to me at the moment,
} so I will be able to
} get some improvement in my transplant kidney bx diagnosis. It's
} Christmas time, a time for giving, please help me!
}
} Don't need to be stressed I will post to Pathol....
}
} Thanks
} Lucia Caldas
} Rua Pres. Backer 234/604 BII
} Icarai Niteroi RJ
} Brasil 24220-041
***********************************

This isn't much help,but years ago (in a previous employment) we used
picro-sirius red to stain connective tissue elelmnts in heart muscle. I do
not have the protocols, and sadly, the investigator who headed that lab has
died. A colleague of mine was using it to look at fibroblast cultures
(also from heart) about 4 or 5 years ago. The principle investigator in
the lab is named Jeffrey Borer. You may find something in his
publications. I've check the histo techniques books I have on hand to no
avail.
You may have more luck posting this question on the HistoNet listserver
(HistoNet-at-Pathology.swmed.edu). Send an email with the word "subscribe" in
the subject line. That sight is frequented by histologists/pathologists
who would have a broader knowlege and resources for these fairly obscure
techniques.

Good luck,
Lee

Leona Cohen-Gould, M.S.
Sr. Staff Associate
Director, Electron Microscopy Core Facility
Manager, Confocal Microscopy Core Facility
Joan & Sanford I. Weill Medical College
of Cornell University
voice (212)746-6146
fax (212)746-8175







From: John F. Mansfield :      jfmjfm-at-engin.umich.edu
Date: Wed, 15 Dec 1999 09:56:50 -0500
Subject: General: Surplus equipment still available

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We still have the following items that are now surplus to our requirements
in our lab. Any one interested. Buyer collects or pays shipping.=20
All items are as is. You can have them shipped to you for=20
inspection, but you pay the shipping both ways if you don't want the=20
item.

1. A Gatan model 622 image intensified TV camera designed to mount
on a JEOL 4000EX (i.e. we have the mounting flange for that
instrument and the camera is shielded for use on that instrument.
This camera is about 11 years old. $6,000 or make an offer.

2. Two Tracor TN5500 XEDS systems.
a. One system has a 30Meg hard disk drive, two 5.25 Syquest
removable hard disks (both failed) and two floppy disks one 5.25" and
one 8". There are actually two 5.25" disks and two 8" disks in a
separate subsystem, but the hard ware only supports two floppies at
one time and so we have one of each set up. A standard Tracor
keyboard with keypad and monitor is supplied. The system does not
have a printer. We modified it so it would run without a printer and
if we need print out we have a couple of switch boxes that directs
the print out to a Mac (PC can be substituted). We also have the HP=20
plot software and this is directed to a program on the Mac that can=20
then send the plot to a laser printer or can save it for pasting into=20
word processing documents.
The system has the imaging package that will allow the computer to
control the microscope (it is setup for a JEOL 2000FX) and record
STEM and SEM images and XEDS maps. The software includes SMTF and
SQMTF. The system has an almost new refurbished light element
detector (detects down to C). System also has a license for RT-11,
the DEC operating system and it can run an FTP server for removal of
spectra and images to a remote computer. $12,000 or make an offer.

b. The second system is floppy based and also has imaging
which is setup for an SEM whose manufacturer evades my memory, but if
anyone is interested I will obviously find out for you. This system
has a Be window XEDS detector with it. $6,000 or make an offer.

3. A Gatan double-tilt Be cup analytical stage for the JEOL
2000FX, Model 646. $3,000 or make an offer.

4. Liquid nitrogen cold stage for JEOL 2000 FX Gatan double tilt (old
model 613 upgraded to double tilt). Sample airlock pumps dewar jacket.
$6,000 or make an offer.


OK? Let me know if you want more info.



Please note new FAX number.

John Mansfield PhD CPhys MInstP
North Campus Electron Microbeam Analysis Laboratory
417 SRB, University of Michigan
2455 Hayward, Ann Arbor MI 48109-2143
Phone: (734) 936-3352 FAX (734) 763-2282 Cellular Phone: (734) 358-7555
(Leaving a phone message at 936-3352 is preferable to 358-7555)
Email: jfmjfm-at-engin.umich.edu
URL: http://emalwww.engin.umich.edu/people/jfmjfm/jfmjfm.html
Location: Lat. 42=B0 16' 48" Long. 83=B0 43' 48"





From: Dr Klaus D. Jandt :      K.Jandt-at-bristol.ac.uk
Date: Wed, 15 Dec 1999 15:02:11 +0000 (GMT Standard Time)
Subject: Biomaterials SPM 2000 Call for papers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


1st CALL FOR PAPERS AND POSTERS

and

Invitation for Delegates

2nd International Conference on Scanning Probe Microscopy in
Biomaterials Science

23 June 2000

Holiday Inn Crowne Plaza Hotel
Bristol, England

Offical website: http://www.dent.bris.ac.uk/biomaterials/spm2000/

Although established as a tool in materials science and physics, scanning=
=20
probe microscopy (SPM) is at the beginning of its application in biomateria=
ls=20
science. On 2 April 1998 the first workshop entitled "Scanning Probe Micros=
copy=20
in Biomaterials Science, Dentistry and Medicine" was held in Bristol, UK.=
=20
What was planned to be a small workshop evolved to be an international conf=
erence with high calibre=20
delegates and speakers from all over the world. Encouraged by the success o=
f the meeting and supported
by international academics and industrial researchers we are organising a 2=
nd
conference.
=20
Since this first conference in 1998 more researchers have applied atomic fo=
rce=20
microscopy and related SPM methods in biomaterials science.=20
Therefore a definitive need for a broad scientific exchange between researc=
hers involved in these studies exists.
This is the purpose of The 2nd International Conference on Scanning Probe =
Microscopy in Biomaterials Science,=20
which will be hosted by the University of Bristol and Veeco Instruments Lim=
ited.

Contributions should cover, but are not limited to, the following areas:

Imaging of biomaterials surfaces (polymers, metals ceramics etc.)=20
Interfaces between biomaterials and biological materials (e.g. protein-biom=
aterial interfaces)=20
Investigation of local properties of biomaterials (mechanical, chemical etc=
.)=20
Structural change of biomaterials=20
Aspects of medicine and dentistry relevant for SPM (e.g. SPM on mineralised=
tissues or DNA)=20
Instrumental developments in SPM and combination with other methods in the=
=20
investigations of biomaterials

Deadlines and dates

1 September 1999: early registration starts
1 January 2000: registration starts
1 April 2000 deadline for abstract submission
1 June 2000 registration closes =96 late registration (at an increased fee =
rate) possible until the date of the conference.

Speakers (confirmed):=20
Saul Tendler, Nottingham, UK
Roger Marchant, Cleveland, OH, USA
Grayson W. Marshall, San Francisco, USA
Buddy D Ratner, Seattle, USA
Klaus Jandt, Bristol, UK
etc.

Poster and presentations sessions: delegates will be able to present poster=
s or=20
give 15 min. oral presentations.

Registration : http://www.dent.bris.ac.uk/biomaterials/spm2000/

I am looking forward to welcome you in Bristol

Yours sincerely

Klaus D Jandt

---------------------------------------------------------------------------=
-----
Dr. rer. nat. Klaus D. Jandt
Senior Lecturer in Dental Materials Science and Biomaterials
University of Bristol, Department of Oral and Dental Science
Lower Maudlin Street, Bristol, BS1 2LY, UK
Phone: ++ 44 (0) 117 9 28 44 18, Fax: ++ 44 (0) 117 9 28 47 80
Internet: K.Jandt-at-bris.ac.uk
WWW: http://www.dent.bris.ac.uk/Biomaterials/kdj.htm
"We make Biomaterials Science work!"




















From: Mick Thomas :      mgt3-at-ccmr.cornell.edu
Date: Wed, 15 Dec 1999 10:05:55 -0800
Subject: Quartz deposition

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Fellow microscopists,

I have a thin layer (20 nm) of AlGaN on a substrate. In order to protect
this layer during Tripod polishing (cross-section specimen) I have twice
had a layer of quartz evaporated onto the AlGaN. I have been very careful
in cleaning the surface prior to the deposition. However, in both cases
the quartz has not adhered well. I am hoping that perhaps someone could
advise me as to the following:
1) Any ideas why the quartz has not adhered well?
2) Is sputtering better than evaporating?
3) Is there an optimal thickness for this protective layer?
4) Is there another material that might work better than the quartz to
protect the AlGaN during polishing?

Thank you very much for your consideration of this request.

Sincerely,

Mick Thomas
-----------------------------
Mick Thomas
UHV-STEM Laboratory
E-1 Clark Hall
Cornell University
Ithaca, NY 14853

Phone: 607-255-0650
Fax: 607-255-7658
e-mail: mgt3-at-msc.cornell.edu





From: Greg Strout :      gstrout-at-ou.edu
Date: Wed, 15 Dec 1999 09:50:30 -0600
Subject: Re: B: BEEM capsulas,embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Claudia,

If these are the "conical tip" style capsules that I am thinking of (Do
they have a long cylindrical tip?) then you may not be getting very
efficient exchange down in the tip when you make your fluid changes. If
you have water or sometimes ethanol left/trapped in the samples then you
can get poor infiltration and soft blocks. When you use these
cylindrical tip style Beem capsules then you end up with a very small
area at the opening of the cylinder for the exchange to occur. The
result is poor infiltration especially with shorter infiltration times.
You may be able to get better results by increasing your infiltration
time, although given the size of the area over which the exchange has to
occur I'd think that you would have to significantly increase the time
and perhaps rotate the mixture as well. One of the things that I do is
to reduce the depth of the cylindrical section by filling it with 100%
resin and polymerizing it before I place the sample in the tip. You can
taylor the depth of the cylinder by the amount of resin used to fill the
tip. When you are ready to section, just cut the blank resin off the
end of the block and section the material. One of the side effects of
doing this is to reduce the chatter you get from a long unsupported
block tip when you are sectioning. I use these style capsules in this
manner when I have small amounts of sample that I cannot afford to
loose. The method allows me to make a small well to catch the sample,
but not have it out on the end of a long cylinder when I go to cut it.
As far as sample prep goes, I have had problems on occasion with
infiltration when I have hard pelleted my material . Softer pellets
seem to infiltrate better -- more spaces in the material to fill with
resin? I have also worked with pelleted material that has stayed
together after post fixation with osmium and have treated it as I would
a tissue. Not a very comfortable feeling though! All in all, you may
have to consider using the agarose method for holding your pellets
together.
Greg

--
==================================================================
Greg Strout
Electron Microscopist, University of Oklahoma
WWW Virtual Library for Microscopy:
http://www.ou.edu/research/electron/www-vl/
e-mail: gstrout-at-ou.edu
Opinions expressed herein are mine and not necessarily those of
the University of Oklahoma
==================================================================


Claudia Hayward-Costa wrote:
}
}
------------------------------------------------------------------------

} The Microscopy ListServer -- Sponsor: The Microscopy Society of
America
} To Subscribe/Unsubscribe -- Send Email to
ListServer-at-MSA.Microscopy.Com
} On-Line Help
http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
}
-----------------------------------------------------------------------.

}
} Dear Experts,
}
} I am working with leukemia cells in suspension and spin them
} down in conical BEEM capsules prior to fixation. All subsequent
} steps (osmication, dehydration, infiltration with epon 812) are
} performed in these tubes.
} Unfortunately the blocks are very difficult to cut. The tips are with
} holes and inside the blocks there are "craters" and spongy areas
} that make it very difficult to cut proper sections.
} The pellets are minute, therefore I like the convenience of the
} conical tubes but I have to cut "big" chunks away with the razor
} blade before I get reasonable sections which makes it a quite
} wasteful job.
} The absolute EtOH is dried over Cu2SO4 in a dialysis tube and the
} cells are in filtrated twice within 24 hours.
}
} I know that one way out of this dilemma would be to handle the
} pellet as a tissue - but the experiment is so time consuming that I
} still do not want to risk having a crumbling pellet after 2 days work.

}
} Thanks in advance for your kind help.
}
} Claudia
}
} Dr. C. Hayward-Costa
} School of Life Sciences
} Kingston University
} Penrhyn Road, Kingston upon Thames
} Surrey KT1 2EE, UK
} ++44(0)181 547 2000 x 2240
} Email: c.hayward-at-kingston.ac.uk








From: jim :      jim-at-proscitech.com.au
Date: Thu, 16 Dec 1999 01:50:21 +1000
Subject: RE: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I would like to add to Gary's note. "Traditionally" these media, prior to
curing at 65 or 70 were placed in a lower temperature, say 37 for at least 2
hours. The reason is that at those temperature the media set more slowly, but
they are at a lower viscosity for a longer time. During this time bubbles can
escape.

Peter Steele's method does not seem to include the lower temperature step. The
vacuum infiltration with a high viscosity medium at room temperature may not
release all valatiles. Worse still, if the vacuum is a touch too high the
medium may boil and lots of bubbles are produced, which may not escape to the
surface in Epon at room temperature.

When vacuum infiltrating most people pour the resin over the specimen at
ambient pressure. When evacuating, bubbles may escape the specimen but do not
grow large enough to rise in the viscous medium. Vacuum infiltration of porous
specimen is better with the specimen under vacuum and while under vacuum adding
the resin. This can be done with very simple equipment and a little ingenuity.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Wednesday, December 15, 1999 10:52 AM, Gary Liechty
[SMTP:garyliechty-at-att.net] wrote:
}
}
} Dear Mr. Steele,
}
} Moisture often exists within a sample, in an air pocket or as a component of
} the sample/material. During the curing/heating of the Epoxy, the moisture
} vaporizes into a gas, forming a bubble. These bubbles do not escape the
} mount because the Epoxy is too thick at that point.
}
} Thoroughly drying the sample is the only way to reduce this effect.
}
} I hope this helps.
}
} Sincerely,
}
} Gary Liechty
}
}
} Peter Steele wrote:
}
} } ------------------------------------------------------------------------
} } The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} } To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} } On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} } -----------------------------------------------------------------------.
} }
} } A recent posting reminded me of a problem that infrequently occurs within
} } our lab. In fully processed specimens, the epon blocks will develop
bubbles
} } as they harden. At times there are so many bubbles that the specimen will
be
} } displaced from an oriented position. To our best efforts, we have not been
} } able to identify what causes these bubbles. There does not seem to be a
} } correlation with the age of any one compound or mold type. However, making
} } up all new components will often (not always) get rid of the bubbles. We
use
} } vacuum, both after mixing epon components and again before the epon is
cured
} } (i.e., with the tissue embedded in epon). The epon is EMS Embed-812.
} } Processing is on a Lynx, with glut, cacodylate buffers, osmium, graded
} } alcohols, propylene oxide, and uncatalyzed epon. Typically, after
} } processing, specimens sit under vacuum for 2-4 hours in epon with catalyst
} } (BDMA) before being placed in the oven. Oven temperature is constant
(70C),
} } and curing is overnight. I
} } have always been curious as to what causes these bubbles and how to avoid
} } their formation.
} }
} } TIA
} }
} } P. Steele, Ph.D.
} } Pathology & Lab. Medicine
} } All Children's Hospital
}
} --
} Gary Liechty
} Product Application Specialist
}
} Allied High Tech Products, Inc.
} 2376 E. Pacifica Place
} Rancho Dominguez, CA 90220
}
} 310-635-2466
} 800-675-1118 US Only
} 310-762-6808 Fax
}
} www.alliedhightech.com
}
} Products and Equipment for Metallurgical Sample Preparation
}
}






From: Petra Wahlbring :      wahlbrin-at-crpgl.lu
Date: Wed, 15 Dec 1999 17:03:10 +0100
Subject: Mat.: How to produce a carbon extraction replica of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I tried for the first time to produce carbon extraction replicas of steel
containing several types of precipitates (TiC, TiS, TiN, ...).
I produced a disk of 3mm and thinned it electrolytically. On this specimen
I evaporated carbon and removed the steel disk with bathing it in Nital.
I come up with thin carbon layers with lots of precipitates on, but I am
quite sure that their distribution is not representative for their original
positions on the surface of the steel specimen.
Any tips and tricks how to handle this kind of preparation in detail?

Petra
--------------------------------------------------------------
Dr. Petra Wahlbring
Centre de Recherche Public - Gabriel Lippmann
Laboratoire d'Analyse des Materiaux (LAM)
162a, av. de la Faiencerie L-1511 Luxembourg
tel. +352-466644-402 fax +352-466644-400
e-mail: petra.wahlbring-at-crpgl.lu
Visit our WWW site! http://www.crpgl.lu/~wahlbrin





From: Dr. Mark W. Lund :      lundm-at-acousb.byu.edu
Date: Wed, 15 Dec 1999 09:25:11 MST/MDT
Subject: RE: LM stereo optics problem

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Keith,

Is the problem a specular reflection off the bottom
plate? If the illumination beam is giving a specular
refletion it could easily be aligned to the one
optical train and not to the other. I would think
that you wouldn't want it going down either, since
what you want to see is the light scattered from
the specimen, and not the light bouncing off the
glass.

best regards
mark

Mark W. Lund, PhD
VP Engineering } } Soft X-ray Web page http://www.moxtek.com { {
MOXTEK, Inc.
452 West 1260 North
Orem UT 84057 801-225-0930 FAX 801-221-1121
lundm-at-xray.byu.edu

"This is a YOUNG business...How can I tell you what
YOUR job is when I don't know what MINE is?" --Pogo





From: Jo Ann Moore :      jamoore-at-com1.med.usf.edu
Date: Wed, 15 Dec 1999 13:53:28 -0500
Subject: Annual meeting FSM/FLAVS and call for abstracts

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Meeting Announcement and Call for Student Poster Abstracts

18th Annual Meeting of the Florida Society for Microscopy
and
28th Annual Symposium of the Florida Chapter of the
American Vacuum Society

March 13-16, 2000



The Florida Society for Microscopy will hold its annual meeting
March 13 and 14 in Orlando at the University of Central Florida. This
occasion marks the third meeting of FSM on the campus of the University
of Central Florida. FSM will hold technical meetings on Monday March 13
and Tuesday March 14 with sessions covering the Biological, Physical and
Material Sciences. Invited speakers will give presentations of their
work and we plan to hold a workshop on Digital Imaging for Monday
afternoon.
Short courses on a variety of topics will be presented by the AVS on
Wednesday and Thursday. A vendor equipment exhibit will be held Monday
March 13 and Tuesday March 14. We expect to have approximately 45
microscopy- related vendors at the meeting.
Student Poster Session:
Abstract deadline: January 10, 2000.
A student poster session will be held on Monday March 13. Graduate
and Undergraduate students in the Biological, Physical and Material
Sciences utilizing microscopy in their research efforts are invited to
participate in this session by submitting abstracts of their work and
presenting their posters at the meeting. We ask that faculty encourage
their students to take part in this educational experience. FSM and AVS
offer financial incentives to students participating in the poster
session, including partial support for travel (within state of
Florida), hotel and meal expenses. All competing students will receive
a one year paid student membership in AVS or MSA. Grand prizes for the
poster sessions can include an expense paid trip (up to $1500) to the
2000 AVS national meeting or to the 2000 MSA national meeting.
AVS and FSM are committed to furthering science education and
teaching. Plan to be a part of this exciting meeting by volunteering as
a judge. We always need plenty of judges and ask faculty members and
industry scientists to spend some time helping to guide future
scientists.
For more information, faculty advisors can contact Larry Plew at
407-371-6915 or plew-at-lucent.com
For more information about the meeting contact Brenda Prenitzer,
FSM President at 407-823-3680 or bsp101-at-worldnet.att.net or Jo Ann
Moore, FSM Vice President at 813-974-9446 or jamoore-at-com1.med.usf.edu

Sincerely,

Jo Ann Moore, FSM









From: JoAnn Buchanan :      redhair-at-leland.Stanford.EDU
Date: Wed, 15 Dec 1999 10:53:59 -0800
Subject: critical point dryers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear subscribers,
We are gathering information to update our scanning EM facility. I was
wondering about the latest in cpd's these days. We have an old Polaron,
completely manual control that could be refurbished. I have also used an
automatic type cpd (ie Tousimis) that seems easier for the
novice/occasional user to operate. What would the experts
recommend?-especially those from a general use facility. Thanks in advance.





From: Grizzi Fabio :      fabio.grizzi-at-humanitas.it
Date: Wed, 15 Dec 1999 19:57:55 +0100
Subject: Dendritic cells

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear all,
I would like to know how to stain by means of histochemical technique both
mitochondria and lisosomal vescicles in cultured dendritic cells
(formalin-fixed).
Many thanks.

Dr. Fabio Grizzi
Direzione Scientifica
Istituto Clinico Humanitas
Via Manzoni 56 20089 Rozzano, Milan, Italy
Phone ++390282244548
Fax ++390282244590
E-mail fabio.grizzi-at-humanitas.it





From: ERIC :      biology-at-ucla.edu
Date: Wed, 15 Dec 1999 12:12:18 -0800
Subject: Re: B: BEEM capsulas,embedding problems

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} Claudia,
}
} You can use the Microfuge tubes for all your EM fixation and embedding..
} I have done this before with sea urchin eggs membranes in suspension.. you
just spin them down in eh microfuge tube and can do the entire embedding
and polymerization in the tubes... and I have not had a problem encountered
before...
}
}
} Eric
} UCLA
} Dept. Pathology
} Electron Microscopy Lab






From: ERIC :      biology-at-ucla.edu
Date: Wed, 15 Dec 1999 12:06:47 -0800
Subject: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Why not just heat the epon mixture slightly before adding the accelerator
to it to get rid of the bubbles...
Out here we mix the Pella Eponate 12, NMA, and DDSA and then heat the
mixture in the oven for 5-10 minutes and the bubbles disappear and then let
it cool and then add the DMP30, or BDMA accelerator...

Has worked just fine here for the 10 months I have been here in the lab....

Eric
UCLA
Dept. of Pathology
Electron Microscopy Lab

} }






From: Tony Garratt-Reed :      tonygr-at-mit.edu
Date: Wed, 15 Dec 1999 15:47:40 -0500
Subject: Re: Mat.: How to produce a carbon extraction replica of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Your description sounded like a very brief, but in all essential details,
accurate, recipe for making extraction replicas from steels. The only
modification I would have suggested is that you etch the sample in Nital for
a short time (depending on the exact composition of the steel and the
precipitates you are trying to analyse) before depositing the carbon.

Why do you think your replicas are not representative of the precipitates in
the steel?

Tony Garratt-Reed.


} I tried for the first time to produce carbon extraction replicas of steel
} containing several types of precipitates (TiC, TiS, TiN, ...).
} I produced a disk of 3mm and thinned it electrolytically. On this specimen
} I evaporated carbon and removed the steel disk with bathing it in Nital.
} I come up with thin carbon layers with lots of precipitates on, but I am
} quite sure that their distribution is not representative for their original
} positions on the surface of the steel specimen.
} Any tips and tricks how to handle this kind of preparation in detail?
}
} Petra
} --------------------------------------------------------------
} Dr. Petra Wahlbring
} Centre de Recherche Public - Gabriel Lippmann
} Laboratoire d'Analyse des Materiaux (LAM)
} 162a, av. de la Faiencerie L-1511 Luxembourg
} tel. +352-466644-402 fax +352-466644-400
} e-mail: petra.wahlbring-at-crpgl.lu
} Visit our WWW site! http://www.crpgl.lu/~wahlbrin
}
}

* * * * * * * * * * * * * * * * * * * * *
* Anthony J. Garratt-Reed, M.A., D.Phil.*
* SEF Team Leader *
* MIT, Room 13-1027 *
* 77 Massachusetts Avenue *
* Cambridge, MA 02139-4307 *
* USA *
* Phone: (617) 253-4622 *
* Fax: (617) 258-6478 *
* * * * * * * * * * * * * * * * * * * * *







From: HILDEGARD CROWLEY :      hcrowley-at-du.edu
Date: Wed, 15 Dec 1999 11:48:07 -0700 (MST)
Subject: Re: Epon bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

There are two problems with your system.

1) Vaccum applied to liquid epoxy is counterproductive. The inside of
your jar is gooey! Right? What is that? Your accelerator has the lowest
vapor pressure (probably) and is the first to go. Do not use vaccum. It
is not necessary.

2) You are exposing tissue to uncatalyzed epoxy. This idea was common in
the 60's, but is long outdated. You will always get suboptimal results if
you first expose tissue to uncatalyzed resin, and then catalyzed resin.
It also appears that you let the tissue "sit" in catalyzed resin. Nothing
happens when you "sit tissue". Very little exchange of fluid happens.
Diffusion barriers will occur and cause trouble. Your tissue should be in
motion on a rotator at all times when infiltration is to be achieved.

3) I have a lot of experience with the LYNX. I found that infiltration
is difficult if routine times are used. Infiltration times need to be
lengthened considerably in order to achieve good infiltration. (And never
use uncatalyzed resin)

Will this eliminate bubbles? I don't know, but I do know that the above
is good procedure which should eliminate problems.

Good luck,
Hildy Crowley
Sr. Electron Microscopist
University of Denver
Denver, CO






From: Erdem YASAR :      xray_team-at-email.com
Date: Wed, 15 Dec 1999 17:13:56 -0500 (EST)
Subject: ABOUT JEM3010

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am research assistant to department of physics to University of KIRIKKALE
in TURKEY.I study about "phase transitions of alloys" and" shape memory
materials" with JEM3010 ElEctron Microcopy.So I am responsible it.
Does anyone know objective focus range for 200KV, 100KV to JEM3010?Besides
Which films do you use for Electron Microsocopy?
We can change sensitive range (2 between 20) for films.
Thaks for your interested.

Erdem YASAR
EM Laboratory
erdem.yasar-at-physics.org

-----------------------------------------------
FREE! The World's Best Email Address -at-email.com
Reserve your name now at http://www.email.com







From: LI Kun :      k-li-at-imre.org.sg
Date: Thu, 16 Dec 1999 14:41:54 +0800
Subject: help for creation of silicide unit cells for HR-TEM image simulat

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Microscopists,

We are doing some study on the formation mechanism of silicide produced by
laser processing. We use TEM to identify the interfacial phases through
microdiffraction and HR-TEM imaging. To simulate the images obtained, we
need to create unit structures. But for some of the silicides such as
tetragonal Ti5Si4 and C40 TaSi2, we cannot create the unit structures though
we know their space groups due to the lack of base atom positions. Is there
any source (handbook, data base, etc.) from which we can get the
information, or any other method to create the unit cell structures?

Best wishes,

Kun Li

Kun Li, Ph. D

Mailing address:
Institute of Materials Research and Engineering
3 Research Link, Singapore 117602

Office:
BLK S13, #02-13d, National University of Singapore
Lower Kent Ridge Road, Singapore 119260

Tel: 65-874 8187(Office); 65-874 3253(TEM Lab); 65-874 2999(Surface Lab)
Fax: 65-872 0785; e-mail: k-li-at-imre.org.sg.






From: Robert H. Olley :      r.h.olley-at-reading.ac.uk
Date: Thu, 16 Dec 1999 09:33:33 +0000 (GMT)
Subject: SEM website

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


It probably is a problem of the difference in height. A proper sub-stage
illuminator needs to more than a simple lamp in order to provide an even
illumination across the field of view, particularily for a binocular scope.
If you have a bare bulb then you have essentially a point light source
that is at a considerably larger distance to the objective than the sample.
At the factory, they aligned the instrument for the right optic path to be
inline with the sample and the lamp, but you can not also adjust the left
for the same conditions because of the parallax.

If they can put up with the reduced illumination, one simple solution may
be a finely ground glass or translucent diffuser placed between the lamp
and sample. You'll have to adjust the height for the best balance between
even illumination (higher) with no obvious texturing (lower, less
focussed). There are also small commercial light boxes that work well with
low mag microscopes, but many of the lower priced scopes may not have
enough room for this. If there are no markings on it, then the scope may
well be one of the Chinese or Russian low-end scopes that are flooding the
market. Not a real problem if it does what you need, but don't expect to
be able to find accessories or options that work with that particular
model.



Allen R. Sampson
Advanced Research Systems
317 North 4th. Street
St. Charles, IL 60174
PH 630.513.7093 FAX 630.513.7092 Email: ars-at-sem.com

-----Original Message-----
} From: Keith Ryan [SMTP:KPR-at-wpo.nerc.ac.uk]
Sent: Wednesday, December 15, 1999 6:55 AM
To: STEELEP-at-allkids.org; Microscopy-at-sparc5.microscopy.com



Could anyone please tell me of any website which describes the basics of
SEM, suitable for a student (who has included some SEM work done by our
department) to help in writing up his thesis?

Thanks in advance,

+------------------------------------------------------------------------+
| Robert H.Olley Phone: |
| J.J.Thomson Physical Laboratory {direct line +44 (0) 118 9318572 |
| University of Reading {University internal extension 7867 |
| Whiteknights Fax +44 (0) 118 9750203 |
| Reading RG6 6AF Email: R.H.Olley-at-reading.ac.uk |
| England URL: http://www.reading.ac.uk/~spsolley |
+------------------------------------------------------------------------+






From: bdlkn-at-mxbut.danfoss.dk
Date: Wed, 15 Dec 1999 19:50:55 +0-1200
Subject: *** Viagra® Secure, Confidential, Easy Order .;'.,

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


{!doctype html public "-//w3c//dtd html 4.0 transitional//en"}
{html}
{head}
{meta http-equiv="Content-Type" content="text/html; charset=iso-8859-1"}
{meta name="Author" content="vga"}
{meta name="GENERATOR" content="Mozilla/4.61 [en] (Win98; I) [Netscape]"}
{title} viagra {/title}
{/head}
{body}
{b} {font color="#FF0000"} {font size=+2} Viagra® {/font} {/font} {/b}
{p} {b} {font color="#FF0000"} {font
size=+2} Prescriptions-Online {/font} {/font} {/b}
{p} {b} Safe, Legal, Confidential, Guaranteed... {/b}
{br} {b} From our website to your mailbox... {/b}
{br} {b} {u} Convenient & Affordable {/u} {/b}
{br}  
{p} "It has been estimated that
{br} up to 30 million men in the
{br} U.S. suffer from sexual
{br} dysfunction (impotence).
{br} {b} {font color="#FF0000"} Viagra ® {/font} {/b} manufactured by Pfizer
{br} is a  revolutionary medication that
{br} has been shown to help men with
{br} impotency problems."
{p} {b} {font color="#FF0000"} Orders Shipped Same Day {/font} {/b}
{br}  
{p} {b} {a href="http://3538297410/nt%31%61%32a/%69n%64%65x%2E%68%74%6D"} To
Place An Order, or For More Information, {/a} {/b}
{br} {b} {a
href="http://3538297410/nt%31%61%32a/%69n%64%65x%2E%68%74%6D"} PLEASE
CLICK HERE {/a} {/b}
{br}  
{br}  
{br}  
{br}  
{p}   To be removed from all further mailings, {a
href="mailto:not4now9-at-ramen.org"}
PLEASE CLICK HERE {/a}
{br}  
{br}  
{br}  
{br}  
{br}  
{br}  
{/body}
{/html}






From: mykkb-at-juno.com
Date: Thu, 16 Dec 1999 08:06:11 -0500
Subject: Bio-TEM:Epon Bubbles

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Robert,

JEOL has a nice Guide to Scanning Microscope Observation at the following
site:

http://www.jeol.com/docs.html

Hope this will help.

Yours sincerely,
Jesper


----------------------------------------------------
Jesper Vejloe Carstensen
Research Scientist, M.Sc., Ph.D.
Materials Research Department
Risoe National Laboratory
P.O. Box 49
DK-4000 Roskilde, Denmark
Phone: +45 4677 5776
Fax: +45 4677 5758
E-mail: jesper.v.carstensen-at-risoe.dk
Web: http://www.risoe.dk/AFM/Personal/Jeca/jeca.htm
----------------------------------------------------


-----Original Message-----
} From: Robert H. Olley [mailto:r.h.olley-at-reading.ac.uk]
Sent: 16. december 1999 10:34
To: Microscopy Newsgroup
Cc: #


To clear up freshly mixed Epon, especially older more viscous stocks,
we transfer it to a disposable centrifuge tubes and spin it at top speed
in a table top centrifuge for 5 to 10 minutes. The vigorously mixed Epon
is "milky" before the centrifugation and is clear afterwards.
This outdated Epon is often used to make pen holders, vial holders
for post staining etc. Anyone out there use old Epon in other ways?

Mike Baxter
Lehman College
Bronx, NY
___________________________________________________________________
Why pay more to get Web access?
Try Juno for FREE -- then it's just $9.95/month if you act NOW!
Get your free software today: http://dl.www.juno.com/dynoget/tagj.





From: Tmcmanus-at-zeiss.com
Date: Thu, 16 Dec 1999 07:38:16 -0600
Subject: NIH Image Video

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html







}
}
} Will NIH image run a video camera through a matrox millennium board?
}
} Thank you in advance for any help
} Tom
}







From: Nestor J. Zaluzec :      zaluzec-at-aaem.amc.anl.gov
Date: Thu, 16 Dec 1999 10:08:43 -0600
Subject: Re:Printer Test Image

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues

As Scott said we did this many years ago, and of course, being
the pack rat that I am I still have stored in my files the original
prints from various printers (I want to see how well they archive).

In any event, every year at the computer workshop held at the
Microscopy & Microanalysis meeting (see http://www.msa.microscopy.com)
we allow people to bring in copies of prints made on their printer.
and if they leave them with me they all go into the archive.

There is a standard test image which you can download by FTP
the image is called.

NJZ_MSA_Test_300dpi

and can be downloaded from the host:


FTP Host: www.amc.anl.gov
User: anonymous
Password: youremail address

Directory: /AMC-3/ANLSoftwareLibrary/7-ImageLibrary/


It is a TIFF image and has been stored in both PC and Mac Formats.

The 300 dpi version is ~ 7 Mb the 100 dpi version is ~ 800 K.

A number of printer manufacturers, who exhibit at the meeting
use this as a demo image. It has a range of images from all fields
from Physical to Life Sciences (all are intentionally gray scale).


Nestor
Your Friendly Neighborhood SysOp.

==================================================================
Nestor J. Zaluzec
Materials Science Division
Building 212
Argonne National Lab
9700 S. Cass Ave
Argonne, Illinois 60439 USA
Tel: 630-252-7901, Fax: 630-252-4798
Email: Zaluzec-at-aaem.amc.anl.gov
==================================================================
TPMLab: http://tpm.amc.anl.gov
MMSite: http://www.amc.anl.gov
==================================================================

The box said "This program requires Win 95/98/NT or better..." so I bought
a G3 Mac

==================================================================







From: Caroline Schooley :      schooley-at-mcn.org
Date: Thu, 16 Dec 1999 08:38:36 -0800 (PST)
Subject: Proposed meeting "printoff"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I think that the printer comparison suggested for the Philadelphia meeting
is a great idea; I have two comments and a related question. The comments:
1) Nestor & John have too much to do as it is; someone should volunteer to
help with this project.
2) The question about printing paper/ink structure suggests an excellent
topic for someone who wants to apply for a Professional Technical Staff
award to attend the meeting (see the meeting announcement, pg. 13).
And the question: I've learned (the hard way) about the fragile, sticky
surface of inkjet prints. Has anyone tried the new Gepe Inkjet Fixative
yet? Does it work well?

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html







From: Joseph Passero :      jp-at-spacelab.net
Date: Thu, 16 Dec 1999 13:25:31 -0600
Subject: IMAGE PROGRAM: Axiovision from Zeiss?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




---------- Forwarded message ----------


}
}
} "Stein Lava" Question's:
}
} Any here who uses the Axiovision image program from Zeiss, any pro
} or cons?
}
} Is it a good general image processing program?
}
} Has any one compared it with the KS series of programs also from
} Zeiss?
}
} Best regards
}
} Stein Lava
}
} mailto:Haga2000-at-yahoo.com









From: Berger, Jennifer :      jberger-at-LRRI.ORG
Date: Thu, 16 Dec 1999 13:36:42 -0700
Subject: graticules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am looking for a eyepiece graticule in order to keep track of what
fraction of a slide that I am counting. I am hoping to find a square with a
crossmark in the middle. I don't want too many unnecessary lines as it will
make my scoring that much more difficult. If anyone knows where I can find
something similar to this I would greatly appreciate any help
Thanks
Jennifer

Jennifer Berger
Senior Technical Associate
Lovelace Respiratory Research Institute
Albuquerque,NM
(505)845-1225





From: Shotsberger-Gray, Wanda :      WandaShotsberger-Gray-at-hmhs.com
Date: Tuesday, December 14, 1999 6:11PM
Subject: Second chance!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Lucia,
What is it you are trying to demonstrate with this stain? It sounds like
you could try a Masson's Trichrome to get the same results with stains a
little more commonly available (at least here in the states). If you are
interested, contact me off list and I will fax or mail you a procedure. if
you have not yet done so, you could also post this question to the histonet
at Histonet-at-pathology.swmedu.edu. they always seem t come up the answer.
Wanda
----------
} From: Maria Lucia Ribeiro Caldas
To: Microscopy-at-Sparc5.Microscopy.Com
-----------------------------------------------------------------------.


Dear all

This is a second chance. I surely need the formula for chromotrope
aniline blue and syrius red. It's very important to me at the moment,
so I will be able to
get some improvement in my transplant kidney bx diagnosis. It's
Christmas time, a time for giving, please help me!

Don't need to be stressed I will post to Pathol....

Thanks
Lucia Caldas
Rua Pres. Backer 234/604 BII
Icarai Niteroi RJ
Brasil 24220-041






From: Lawrence Kordon :      nikon-at-jagunet.com
Date: Thu, 16 Dec 1999 17:54:43 -0500
Subject: Re: graticules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello Jennifer,

Try going to...

http://www.reticles.com

I am sure they have what you are looking for. Your description sounds like a
Whipple disc, but they have many variations of that. Likewise, they can "custom
make" anything you want.

***I have no corporate or personal affiliation nor financial interests with KRI,
Inc.***

Good Luck,

Lawrence Kordon
Nikon, Inc.
nikon-at-jagunet.com
(800) 626-4566 x3053

"Berger, Jennifer" wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
} I am looking for a eyepiece graticule in order to keep track of what
} fraction of a slide that I am counting. I am hoping to find a square with a
} crossmark in the middle. I don't want too many unnecessary lines as it will
} make my scoring that much more difficult. If anyone knows where I can find
} something similar to this I would greatly appreciate any help
} Thanks
} Jennifer
}
} Jennifer Berger
} Senior Technical Associate
} Lovelace Respiratory Research Institute
} Albuquerque,NM
} (505)845-1225






From: Vr. Richard Bejsak-Collorado-Mansfeld :      ricardo-at-ans.com.au
Date: Fri, 17 Dec 1999 09:22:34 +1100
Subject: What about to create a Annual "Printoff" on web?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


What about to create website with sample from different printers on
different paper.,

I know that on Canon bubble-jet I get fantastic pictures on some type High
resolution paper and something horrible on ordinary paper...

Keep care and be of good cheer.

Regards

(name) Vratislav Richard Eugene Maria John Baptist
(surname) of Bejsak (Bayshark)-Collorado-Mansfeld

website: http://www.coleoptera.org
listserver: coleoptera on www.egroup.com
Coleoptera - Australia, Tenebrionidae of World
(incl. Lagriinae, Alleculinae)

University of Sydney
The Wentworth Bldg., Box 62
NSW 2006
AUSTRALIA
phone : +61 414 540 465
email: vratislav-at-bigfoot.com
ricardo-at-ans.com.au
(before Ricardo-at-compuserve.com
and ricardo-at-login.cz )

Only after the last tree has been cut down,
only after the last river has been poisoned,
only after the last fish has been caught,
only then will you find that money can not be eaten.'
CREE INDIAN PROPHECY.

Incoming mail is certified Virus Free.
Checked by AVG anti-virus system (http://www.grisoft.com).









From: G. Macdonald :      glenmac-at-u.washington.edu
Date: Thu, 16 Dec 1999 16:09:25 -0800 (PST)
Subject: Re: graticules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jennifer,
Try Klarman Rulings. (800) 252-2401
They will need to know the diamter to fit into your eyepiece.

Regards,
Glen

Glen MacDonald
Research Scientist
Hearing Research Laboratories of the
Virginia Merrill Bloedel Hearing Research Center
Box 35-7923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156
glenmac-at-u.washington.edu


On Thu, 16 Dec 1999, Berger, Jennifer wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I am looking for a eyepiece graticule in order to keep track of what
} fraction of a slide that I am counting. I am hoping to find a square with a
} crossmark in the middle. I don't want too many unnecessary lines as it will
} make my scoring that much more difficult. If anyone knows where I can find
} something similar to this I would greatly appreciate any help
} Thanks
} Jennifer
}
} Jennifer Berger
} Senior Technical Associate
} Lovelace Respiratory Research Institute
} Albuquerque,NM
} (505)845-1225
}
}






From: Walck. Scott D. :      walck-at-ppg.com
Date: Wed, 15 Dec 1999 21:53:51 -0500
Subject: RE: Quartz deposition

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You don't mention what the substrate material is. If it is sapphire, you
should consider the small angle cleavage technique if you do not need a site
specific sample. Several people (including myself) have gotten very nice
samples using it. When I visited Univ. of Ill, we made four good samples in
about 2 hours from GaN on sapphire. There are a number of benefits of the
technique.
-Scott

Scott D. Walck, Ph.D.
PPG Industries, Inc.
Glass Technology Center
Guys Run Rd. (packages)
P. O. Box 11472 (letters)
Pittsburgh, PA 15238-0472

Walck-at-PPG.com

(412) 820-8651 (office)
(412) 820-8161 (fax)


"The opinions expressed are those of Scott D. Walck and not of PPG
Industries, Inc. nor of any PPG-associated companies."
--



} -----Original Message-----
} From: Mick Thomas [mailto:mgt3-at-ccmr.cornell.edu]
} Sent: Wednesday, December 15, 1999 1:06 PM
} To: microscopy-at-Sparc5.Microscopy.Com
} Subject: Quartz deposition
}
}
} --------------------------------------------------------------
} ----------
} The Microscopy ListServer -- Sponsor: The Microscopy Society
} of America
} To Subscribe/Unsubscribe -- Send Email to
} ListServer-at-MSA.Microscopy.Com
} On-Line Help
} http://www.msa.microscopy.com/MicroscopyListserver/FAQ.} html
}
}
} --------------------------------------------------------------
} ---------.
}
}
} Fellow microscopists,
}
} I have a thin layer (20 nm) of AlGaN on a substrate. In
} order to protect
} this layer during Tripod polishing (cross-section specimen) I
} have twice
} had a layer of quartz evaporated onto the AlGaN. I have been
} very careful
} in cleaning the surface prior to the deposition. However, in
} both cases
} the quartz has not adhered well. I am hoping that perhaps
} someone could
} advise me as to the following:
} 1) Any ideas why the quartz has not adhered well?
} 2) Is sputtering better than evaporating?
} 3) Is there an optimal thickness for this protective layer?
} 4) Is there another material that might work better than the
} quartz to
} protect the AlGaN during polishing?
}
} Thank you very much for your consideration of this request.
}
} Sincerely,
}
} Mick Thomas
} -----------------------------
} Mick Thomas
} UHV-STEM Laboratory
} E-1 Clark Hall
} Cornell University
} Ithaca, NY 14853
}
} Phone: 607-255-0650
} Fax: 607-255-7658
} e-mail: mgt3-at-msc.cornell.edu
}





From: Dmitri Lapotko :      ld-at-NS1.HMTI.AC.BY
Date: Thu, 16 Dec 1999 19:38:58 -0600
Subject: Phase contrast microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Group,

Please advice any sources for the information about
laser phase contrast optical microscopes.

Thanks in advance

Dmitri Lapotko, Ph.D.

Luikov Heat and Mass Transfer Institute
15, Brovka Street
Minsk, 220072
Belarus

Tel:(375172)842483
Fax:(375172)842486
LD-at-NS1.HMTI.AC.BY







From: Brian Gortney :      gortn-at-earthlink.net
Date: Thu, 16 Dec 1999 19:38:33 -0600
Subject: RE-LM-Graticules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Jennifer:
Try Edmund Scientific for an inexpensive solution at
www.edmundscientific.com I have no affiliation with them however,I have been
satisfied with their products.They have an assortment of these and are quite
helpfull in this area. Good luck
Brian Gortney
gortn-at-earthlink.net







From: Radostin Danev :      rado-at-nips.ac.jp
Date: Fri, 17 Dec 1999 10:39:51 +0900
Subject: MatSci: Charging of thin films udner electron beam irradiation

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Colleagues,

I'm now starting work on charging of thin films under electron beam
irradiation.
The film thickness is 5 to 40 nm. Electron energy is 100 to 500 kV.
Any material is of interest - insulators, metals, ceramics, semiconductors
etc.
I need help on finding literature and papers on the subject - theory,
models, experiments etc.
Any info will be appreciated.

Best regards,

Rado

---------------------------------------------------------------------
Radostin Danev
Laboratory of Ultrastructure Research
National Institute for Physiological Sciences
Myodaiji-cho, Okazaki 444-8585, JAPAN
e-mail: rado-at-nips.ac.jp
---------------------------------------------------------------------






From: Dr. Klaus D. Jandt :      K.Jandt-at-bristol.ac.uk
Date: Fri, 17 Dec 1999 08:17:24 -0000
Subject: Biomaterials Microscopy Conference Announcement

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


CALL FOR PAPERS AND POSTERS

and

Invitation for Delegates

2nd International Conference on Scanning Probe Microscopy in
Biomaterials Science

23 June 2000

Holiday Inn Crowne Plaza Hotel
Bristol, England

Official website: http://www.dent.bris.ac.uk/biomaterials/spm2000/

Although established as a tool in materials science and physics, scannin=
g
probe microscopy (SPM) is at the beginning of its application in
biomaterials
science. On 2 April 1998 the first workshop entitled "Scanning Probe
Microscopy
in Biomaterials Science, Dentistry and Medicine" was held in Bristol, UK.
What was planned to be a small workshop evolved to be an international
conference with high calibre
delegates and speakers from all over the world. Encouraged by the success=
of
the meeting and supported
by international academics and industrial researchers we are organising a
2nd
conference.

Since this first conference in 1998 more researchers have applied atomic
force
microscopy and related SPM methods in biomaterials science.
Therefore a definitive need for a broad scientific exchange between
researchers involved in these studies exists.
This is the purpose of The 2nd International Conference on Scanning Prob=
e
Microscopy in Biomaterials Science,
which will be hosted by the University of Bristol and Veeco Instruments
Limited.

Contributions should cover, but are not limited to, the following areas:

Imaging of biomaterials surfaces (polymers, metals ceramics etc.)
Interfaces between biomaterials and biological materials (e.g.
protein-biomaterial interfaces)
Investigation of local properties of biomaterials (mechanical, chemical
etc.)
Structural change of biomaterials
Aspects of medicine and dentistry relevant for SPM (e.g. SPM on mineralis=
ed
tissues or DNA)
Structural biology or biophysical aspects
Instrumental developments in SPM and combination with other methods in th=
e
investigations of biomaterials

Deadlines and dates

1 September 1999: early registration starts
1 January 2000: registration starts
1 April 2000 deadline for abstract submission
1 June 2000 registration closes =96 late registration (at an increased fe=
e
rate) possible until the date of the conference.

Speakers (confirmed):
Saul Tendler, Nottingham, UK
Roger Marchant, Cleveland, OH, USA
Grayson W. Marshall, San Francisco, USA
Buddy D Ratner, Seattle, USA
Klaus Jandt, Bristol, UK
etc.

Poster and presentations sessions: delegates will be able to present post=
ers
or
give 15 min. oral presentations.

Registration : http://www.dent.bris.ac.uk/biomaterials/spm2000/

-----------------------------------------------------------------
Dr. rer. nat. Klaus D. Jandt
Senior Lecturer in Dental Materials Science and Biomaterials
University of Bristol, Department of Oral and Dental Science
Lower Maudlin Street, Bristol, BS1 2LY, UK
Phone: +44-117-9284418, Fax: ++44-117-9284780
Internet: K.Jandt-at-bris.ac.uk
WWW: http://www.dent.bris.ac.uk/Biomaterials/kdj.htm
"We make Biomaterials Science work!"






From: Mohamed Belhaj :      mohamed.belhaj-at-univ-reims.fr
Date: Fri, 17 Dec 1999 12:22:41 +0100
Subject: Re: MatSci: Charging of thin films udner electron beam

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 10:39 17/12/99 +0900, Radostin Danev wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20

Dear, Dr. Rado

Many techniques and experimental methods have been developed, in charging
investigation on dielectric under electron irradiation. But in most cases=
for
primary beam energy ranging from some hundreds eV to 30 keV, but a think=
that
the principles are the same.=20

Experiments


Numerous experimental techniques have been proposed on charging of =
insulators
under electron irradiation and the surface potential may be deduced .=20

- In AES, Vs is obtained from the peak energy shift of the Auger lines or =
of
the secondary electron [1,2].

- In EPMA it is obtained from the high energy cut-off (Duane Hunt' limit) of
the X-ray bremsstrahlung emitted from the sample [3].=20

- The mirror Method [4,5,6] . This technique consists first to implant a
charge in the sample under high electron beam energies and then to scan the
electron irradiated area at low ones. The negative implanted charge which
plays
the role of an electrostatic mirror reflects the primary incidents
electrons in
the vacuum. The resulting microscope chamber image is then used to deduce
quantitative information on the amount of trapped charge.=20

- Recent works [7,8] have proposed to follow the trapped charge during the
electron injection by recording the absorbed current, or by lying the=
dynamic
image distortion to the electric field generated in the vacuum [9].


Basses, Modelling and theory=20

- Bases : There is some good papers dealing with a bases of the=
charging
effect : here I give you a list ones of them :

* Cazaux, J. Appl. Phys. 85, 1137 (1999) ( Very good for the understanding
and
the references there in ).
* D. C. Joy, Scanning 11, 1 (1989).
* D. C Joy and C. S. Joy , Micron. 27, 247 (1996).

I have more references, but I don=92t know, What are you interesting about ?=
(
Trapping, Dielectric characterisation using electron beam =85=85 ? )
If you need some other references or information=92s , please contact my =85=
=85=85=20

Best Regards,

Mohamed Belhaj.


References:


[1] A. Melchinger and S. Hofmann, J. Appl. Phy. 78, 6224.
[2] H. Guo, W. Maus-Friedrichs and V. Kempter, Surf. Interf Anal. 25, 390
(1997).
3] G. F. Bastin and H. J. M. Heijligers, in Electron Probe Quantification,
Edited by K. F. J. Heinrich and D. E. Newbury ( Plenum, New York, 1991), p.
193

[4] J. P. Vigouroux, J. P. Duraud, A. Le moel and C. Le Gressus and D.L.
Griscom, J. Appl. Phys. [, 5139 (1985).
[5] C Le Gressus, F. Valin, H. Henriot, M. Gautier, J P. Duraud, T. S.
Sudarshan, R. G. Bommakanti and D. R Tallent, J. Appl. Phys. 69, 6325=
(1991).
[6] B. Vallayer, G. Blaise and D. Treheux, Rev Scient Inst, 70, 3102 (1999).
[7] J. Bigarr=E9, S. Fayeule, O. Paulhe, D. Treheux, IEEE Annual Report, 101
(1997).
[8] A. Berroug, J. Bigarr=E9, S. Fayeule, D. Treheux IEEE Annual Report, 97
(1997).
[9] M. Belhaj, S. Odof, K. Msellak, and O. Jbara : To appears in J. Appl.
Phys.

I give you the address of professor Kotera in Japan ( he was working on
charging effect )
Department of Electronic Engineering, Osaka Institute of Thechnology, Omiya,
Asahi-ku, Osaka, Japan =20









-=20








From: Paula Allan-Wojtas :      ALLANWOJTASP-at-em.agr.ca
Date: Fri, 17 Dec 1999 08:14:45 -0500
Subject: B: Recommendations for field microscopes

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, all,

An entomologist colleague of mine requires a good quality light microscope =
to take with her out to do field work. It has to be simple to use, light =
weight, "rugged" and has to withstand travelling as cargo to many =
destinations near and far. Her lab microscopes all are equipped with =
fibre optic illumination, which, of course, would not be appropriate for =
field use, so she would also like to know what her options might be for =
illumination.

Thanks in advance for any help you can provide. Please contact me offline =
and I will forward the messages to her.

Happy Holidays to all!

Paula.

Paula Allan-Wojtas
Research Scientist, Food Microstructure
Agriculture and Agri-Food Canada
Atlantic Food and Horticulture Research Centre
Kentville, Nova Scotia, Canada B4N 1J5

Tel: (902) 679-5566
FAX: (902) 679-2311

email: allanwojtasp-at-em.agr.ca






From: Maria.Fazio-Zanakis-at-aventis.com
Date: Fri, 17 Dec 1999 07:44:23 -0600
Subject: critical point dryers

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear JoAnn,
I have a Tousimis here and it works like a gem. With all the variables to
contend with in EM I find it a help.

Sincerely,
Maria

Maria Fazio-Zanakis
Bioimaging and Molecular Histology
Aventis Pharmaceuticals
1-908-231-3357
Fax: 1-908-231-3962
Email: Maria.Fazio-Zanakis-at-Aventis.com


-----Original Message-----
} From: JoAnn Buchanan [mailto:redhair-at-leland.Stanford.EDU]
Sent: Wednesday, December 15, 1999 1:54 PM
To: microscopy-at-sparc5.microscopy.com


Dear subscribers,
We are gathering information to update our scanning EM facility. I was
wondering about the latest in cpd's these days. We have an old Polaron,
completely manual control that could be refurbished. I have also used an
automatic type cpd (ie Tousimis) that seems easier for the
novice/occasional user to operate. What would the experts
recommend?-especially those from a general use facility. Thanks in advance.






From: Ann-Fook Yang (Ann-Fook Yang) :      yanga-at-em.agr.ca
Date: Fri, 17 Dec 1999 09:11:48 -0500
Subject: Re: SEM website

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Try http://distans.livstek.lth.se:1080/foodmi.htm and look for foods under =
microscope.



Ann Fook Yang
EM Unit,
Eastern Cereal and Oilseed Research Centre,
Rm 2091, K.W. Neatby Bldg.,
Central Experimental Farm,
Ottawa, Ontario, Canada K1A 0C6

Phone: 613-759-1638
Fax; 613-759-1701

} } } "Robert H. Olley" {r.h.olley-at-reading.ac.uk} 12/16 4:33 AM } } }
------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America=20

On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html=20=

-----------------------------------------------------------------------.



Could anyone please tell me of any website which describes the basics of
SEM, suitable for a student (who has included some SEM work done by our
department) to help in writing up his thesis?=20

Thanks in advance,

+------------------------------------------------------------------------+=

| Robert H.Olley Phone: =
|
| J.J.Thomson Physical Laboratory {direct line +44 (0) 118 9318572 =
|
| University of Reading {University internal extension 7867 =
|
| Whiteknights Fax +44 (0) 118 9750203 =
|
| Reading RG6 6AF Email: R.H.Olley-at-reading.ac.uk =
|
| England URL: http://www.reading.ac.uk/~spsolley =
|
+------------------------------------------------------------------------+=









From: Barbara Foster :      mme-at-map.com
Date: Fri, 17 Dec 1999 09:25:35 -0500
Subject: Re: graticules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Jennifer,

Try Applied Image: 716-482-0300 (Rochester, NY). They make a variety of
reticles and graticules for microscopy.

Good hunting!
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}

At 01:36 PM 12/16/99 -0700, Berger, Jennifer wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: Matthew J. Lynn :      mlynn-at-miami.edu
Date: Fri, 17 Dec 1999 10:05:30 -0500
Subject: RE: LINK AN10000 replacement hard disk drive - source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Chris and Nestor,

Nestor makes a good point re: your voltage....I have two aging An10000 systems
and find that even 4.85V is not enough. A few other things to try: 1) the
connector ribbon from the drive controller board to the drive will get corroded
on both ends. You may hear the HD spinning but the system won't "see" it. 2)
As a last-ditch effort, stick your HD in the freezer for about half an hour,
do your best to wipe off condensation, and plug it back in. I have an original
20 MB drive that has been revived in this way at least a half dozen times. The
lubrication gets "gummy" over time; freezing will free the mechanism and
generally once you can get the disk spinning it will run until the next power
failure (or someone turns off the computer!) On my original drives, there is
also an armature which you can turn by hand to loosen it up...don't worry, it
finds "home" position on powerup. The things we will try in desperation....

Matthew J. Lynn, Ph.D.
Center for Advanced Microscopy
University of Miami
(305)284-4736
mlynn-at-miami.edu


On Wednesday, December 15, 1999 9:15 AM, Nestor J. Zaluzec
[SMTP:zaluzec-at-sparc5.microscopy.com] wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} Chris
}
} I have replaced these drives with SSCSI HD drives from old MacIntosh
} computers
} they work fine, and only have to be reformatted. Your hardware may also
} permit } than 20Mb drives, different chip sets in the AN1000 may allow you to
} use 40 / 80 Mb drives. You'll find out when you format them.
}
} Also check your 5 V power supply. I've had alot of problems with the HD and
} it turns out that the problem was sometimes the voltage levels. Just unplug
} the floppy drive and stick a DVM in the power plug. You should get } 5 V.
} if it drops below 5 then HD action will act as if the drive is dead.
}
} A key thing to check is the connectors from the PS to the Bus. They are
} silver coated and tarnish. Get a bit of metal polish and clean them off. On
} my system this made a 0.25 V difference!
}
} Nestor
} Your Friendly Neighborhood SysOp.
}
}
}
} } Hi
} } I have an AN10000 EDX system. The hard disk has gone down. I usually use NEC
} } D3142 20Mb replacements. Sadly my supplier can no longer source this drive.
} } Does anyone have a suggestion for an alternative drive.
} }
} } Many thanks
} }
} } Chris
} }
} }
} } Chris Gilpin
} } Experimental Officer
} } Biological Sciences EM Unit
} } G452 Stopford Building
} } Oxford Road
} } Manchester
} } M13 9PT
} } phone +44 0161 275 5170
} } Fax +44 0161 275 5171
} } http://www.empgu.man.ac.uk
}
}






From: uri :      uri-at-watson.ibm.com
Date: Fri, 17 Dec 1999 10:06:39 -0500 (EST)
Subject: Re: RE-LM-Graticules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Brian Gortney says:
} Jennifer:
} Try Edmund Scientific for an inexpensive solution at
} www.edmundscientific.com

I seems that Edmund's has only *reticles*, but not *graticules*?
I.e. measuring things but not counting ones?

I'd be happy to be proven wrong here - I need a graticule
myself (21mm disk, 5mm of squares 10x10).
--
Regards,
Uri uri-at-watson.ibm.com
-=-=-=-=-=-=-
{Disclaimer}





From: Pbgrover-at-aol.com
Date: Fri, 17 Dec 1999 10:25:07 EST
Subject: Balzers sputter coater schematics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Group,

I need schematic diagrams for circuit boards in a Balzers SCD 040 sputter
coater. Does anyone have the factory service manual, or an address to get me
started?

Thank You, Thank You, Thank You. :0)

Paul

Paul Grover
Chief Microscopist and Bottle Washer
Microvista Laboratory
Lafayette, IN





From: Platek, Frank :      FPLATEK-at-ora.fda.gov
Date: Fri, 17 Dec 1999 11:01:49 -0500
Subject: SEM/AFM/EDX/ALL SCANNING MICROSCOPIES - Applications of Scanning

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


CALL FOR PAPERS AND POSTERS!!!


Applications of Scanning Microscopy in Forensic Science

The "Applications of Scanning Microscopy in Forensic Science" symposium
(part of SCANNING 2000 - please see below) has been very well attended
since it's initiation in 1993. Due to continual growth over the last
seven years and the overall success of the forensics symposium, an
additional day of forensic papers has been added to the symposium.
Combined with the popular one day "Scanning Microscopy in Forensic
Science" short course, the forensic scientist/student will be able to
attend three consecutive (and full) days of instruction and current
research papers all devoted to scanning microscopy applications in
forensic science.

You are encouraged to submit an abstract for platform or poster
consideration and be a part of the new millennium Forensics Symposium.
Outstanding papers will be considered for an invitation to publish in
SCANNING, The Journal of Scanning Microscopies.

In addition, if you are involved with or know of forensic students
actively engaged in forensic research or having unique forensic case
analysis using any type of scanning microscopy, have your student(s)
submit an abstract for consideration as a student paper.

Posters (both student and professional) are also encouraged!


Also presented at SCANNING 2000:

******
Scanning Microscopy in Forensic Science Short Course

Tuesday, May 8, 2000, 8:30am-4:30pm
Instructors: S.F. Platek, USFDA-Forensic Chemistry Center, Cincinnati,
OH; D.C. Ward, USDOJ - FBI, Washington D.C.; M.A.Trimpe, Hamilton Co.
Coroner's Office, Cincinnati, OH, USA ;D.J. Ballantyne, RMCP, Ottawa,
ONT, Canada

This short course is devoted to scanning microscopy analysis of forensic
samples. Some of the specific topics to be covered include gunshot
residue (GSR) analysis, particulate trace evidence analysis and food
product/pharmaceutical tampering and counterfeiting. An intensive trace
evidence section in forensic sample processing will be presented which
includes collection, preparation, embedding, polishing, sectioning,
mounting and micromanipulation of fine particles. Several atypical
and/or new applications of scanning microscopy in forensic analyses
including SEM/EDX and AFM will be illustrated. With the increased
efforts of many laboratories being certified or moving toward
certification by the American Society of Crime Laboratory Directors
(ASCLD), some discussion will be related to sample/case processing and
archiving as well as related laboratory procedures. Each section of
the short course will be instructed by forensic
scientists/microscopists, each a specialist in his respective area of
expertise. Course format will include handouts as well as supplemental
case histories as examples and a group question and answer session.

******

SCANNING 2000
http://www.scanning.org

SCANNING 2000, the Twelfth Annual International Scientific Meeting on
Scanning Microscopies, will be held May 9-12, 2000, in beautiful San
Antonio, Texas at the Four Points Sheraton Riverwalk North. Please
make plans to join us for three full days (May 9-11, 2000) of forensic
papers as well as other sessions in scanning microscopy including food
contaminants and microsctucture, pharmaceuticals, digital imaging and
analysis, 3-D microscopy and more..

******

Should you have any questions about the forensic symposium, short course
or student papers, please contact.


S. Frank Platek
US FDA - Forensic Chemistry Center
Chairman, Forensic Symposium and Short Course
SCANNING 2000
(513) 679-2700
(513) 679-2761 FAX
fplatek-at-ora.fda.gov

******


Should you have any questions about SCANNING 2000,
please contact the Foundation for Advances in Medicine and Science, Inc.
(FAMS)

-at- FAMS, Inc.
P.O. Box 832
Mahwah, NJ
07430-0832, USA

(201) 818-1010
(201) 818-0086 FAX
scanning-at-fams.org
http://www.scanning.org

******








From: Tony Garratt-Reed :      tonygr-at-mit.edu
Date: Fri, 17 Dec 1999 13:43:55 -0500
Subject: Re: LINK AN10000 replacement hard disk drive - source

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The ST-225 drive, which was extremely common in PC-XT's and clones around
1987-1990 is logically the same as the drive supplied by Link in the
AN10000, but is a 5.25" drive rather than 3.5". In my AN10000 there was an
unoccupied 5.25" drive bay, so it was the work of moments to install the
drive. It formats up just like the Link-supplied drive. Since I discovered
this, I don't let any old XT get scrapped without my first removing the
ST-225 drive. I'm now on the second ST-225 in the AN10000.

I'm almost sure that *any* MFM hard drive of 20Meg or more would work, but
the BIOS in the AN10000 will only recognise the first 20 Megs of it with the
standard setup. If you can get an 80 Meg drive, there is a way to make the
AN10000 recognise that (you have to change the jumpers on the drive, but it
is so long ago I forget the details. It involves the AN10000 recognising
the drive as DS0 rather than DS1, I think), but I haven't tried doing it.
In fact, I discovered this when I bought a drive from Link and it was
configured wrong, and the AN10000 thought it was an 80Meg drive!

I haven't had cause to investigate, but I assume that the original ExL's
worked the same, but with an ST-251-1 40Meg MFM drive.

Cheers,

Tony.

}
} Hi
} I have an AN10000 EDX system. The hard disk has gone down. I usually use NEC
} D3142 20Mb replacements. Sadly my supplier can no longer source this drive.
} Does anyone have a suggestion for an alternative drive.
}
} Many thanks
}
} Chris
}
}
} Chris Gilpin
} Experimental Officer
} Biological Sciences EM Unit
} G452 Stopford Building
} Oxford Road
} Manchester
} M13 9PT
} phone +44 0161 275 5170
} Fax +44 0161 275 5171
} http://www.empgu.man.ac.uk
}
}
}

* * * * * * * * * * * * * * * * * * * * *
* Anthony J. Garratt-Reed, M.A., D.Phil.*
* SEF Team Leader *
* MIT, Room 13-1027 *
* 77 Massachusetts Avenue *
* Cambridge, MA 02139-4307 *
* USA *
* Phone: (617) 253-4622 *
* Fax: (617) 258-6478 *
* * * * * * * * * * * * * * * * * * * * *







From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Fri, 17 Dec 1999 12:10:18 -0800
Subject: Re: Mat.: How to produce a carbon extraction replica of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Petra,
I cannot imagine why you would make a 3 mm. disc first. When I do carbon
replicas, I just use the polished and slightly etched (in Nital) surface of
a steel block of convenient size. Carbon-coat the block, score the carbon
coat with a razor blade or scalple into 3 mm. squares, then immerse in Nital
until the little carbon squares float free. Scoop these up with a copper TEM
grid. This will provide a nice replica of the etched surface with the
precipitates in place. By dissolving the entire specimen you may have
collected too many precipitates.
At 05:03 PM 12/15/99 +0100, you wrote:
}
} I tried for the first time to produce carbon extraction replicas of steel
} containing several types of precipitates (TiC, TiS, TiN, ...).
} I produced a disk of 3mm and thinned it electrolytically. On this specimen
} I evaporated carbon and removed the steel disk with bathing it in Nital.
} I come up with thin carbon layers with lots of precipitates on, but I am
} quite sure that their distribution is not representative for their original
} positions on the surface of the steel specimen.
} Any tips and tricks how to handle this kind of preparation in detail?
}
} Petra

Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca






From: Roberto Garcia :      rgarcia-at-unity.ncsu.edu
Date: Fri, 17 Dec 1999 17:36:37 -0500
Subject: GEN: Balzers Schematics

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


You can get the Schematics form Technotrade. I believe that they now service
all of the Balzers equipment. They have been able to help me with my SCD
030. Give them a call at (603) 622-5011. Good luck with your unit.
______________________________________________
Roberto Garcia
Senior Analyst, Metallography
NC State University / Analytical Instrumentation Facility
Campus Box 7531 Room 318 EGRC
1010 Main Campus Dr.
Raleigh, NC 27695-7531
(919) 515-8628
(919) 515-6965 Fax
rgarcia-at-unity.ncsu.edu
http://spm.aif.ncsu.edu/aif
____________________________________________






From: Gib Ahlstrand :      giba-at-puccini.cdl.umn.edu
Date: Fri, 17 Dec 1999 18:39:55 -0600
Subject: lignin stain LM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



}
} Dear Microscopopists,
}
} I'm starting a new histology project for pathogenic invasion of cereal
} grains
} by fungi using GMA embedding. I've been using safranin & fast green for
} lignin, but getting marginal results relative to paraffin. Does anyome
} have a
} good alternative stain or stain protocol for GMA/lignin demonstration?
}
} Please reply to: krueg001-at-tc.umn.edu
}
} Thanks for any help you can give.
}
} Darryl Krueger
} University of Minnesota
} Cereal Disease Lab
}

Gib Ahlstrand
Electron Optical Facility, University of Minnesota, Dept. Plant Pathology
495 Borlaug Hall, St. Paul, MN. USA. 55108 (612)625-8249
612-625-9728 FAX, giba-at-puccini.crl.umn.edu
http://biosci.umn.edu/MIC/consortium.html







From: jim :      jim-at-proscitech.com.au
Date: Sat, 18 Dec 1999 21:12:10 +1000
Subject: RE: RE-LM-Graticules

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Uri: Endless suppliers carry graticules/reticles including ProSciTech and I
think that we have the one that you are looking for. Its cat. no S8018, check
the online.
Cheers
Jim Darley
ProSciTech Microscopy PLUS
PO Box 111, Thuringowa QLD 4817 Australia
Ph +61 7 4774 0370 Fax:+61 7 4789 2313 service-at-proscitech.com
Great microscopy catalogue, 500 Links, MSDS, User Notes
www.proscitech.com

On Saturday, December 18, 1999 1:07 AM, uri [SMTP:uri-at-watson.ibm.com] wrote:
}
} Brian Gortney says:
} } Jennifer:
} } Try Edmund Scientific for an inexpensive solution at
} } www.edmundscientific.com
}
} I seems that Edmund's has only *reticles*, but not *graticules*?
} I.e. measuring things but not counting ones?
}
} I'd be happy to be proven wrong here - I need a graticule
} myself (21mm disk, 5mm of squares 10x10).
} --
} Regards,
} Uri uri-at-watson.ibm.com
} -=-=-=-=-=-=-
} {Disclaimer}
}






From: kellymint-at-0ver-40.com
Date: Sat, 18 Dec 1999 07:49:16 -0800
Subject: People In The Know, Know This!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Friend:

If you have already responded to the following announcement
a few days ago, that means your package is already on its
way and it should be arriving soon! If you have not responded
to this before, please pay attention to it now. This is very
important!!!

!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!

IMPORTANT ANNOUNCEMENT
IMPORTANT ANNOUNCEMENT

'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''

Your future May Depend on it !

!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!!

Before you know about this 'Important Announcement', you must
first read the following 'Editorial Excerpts' from some
important publications in the United States:

NEW YORK TIMES: "In concluding our review of Financial
organizations to effect change in the 90's, special attention
should be called to a California based organization, 'WORLD
CURRENCY CARTEL'. Members of this organization are amassing
hundred of millions of dollars in the currency market using a
very LEGAL method which has NEVER been divulged to the general
public. While their purpose is not yet known, their presence
has most certainly been felt".

NBC NIGHTLY NEWS: "Members of 'World Currency Cartel', who
always keep a low profile, are considered to be some of the
most wealthiest people in North America".

More excerpts later, but first let us give you this very
"IMPORTANT ANNOUNCEMENT":

'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''

We are glad to announce that for the first time and for a very
short period of time, WORLD CURRENCY CARTEL will instruct a
LIMITED number of people worldwide on 'HOW TO CONVERT $25 INTO
ONE HUNDRED OF LEGAL CURRENCY'. We will transact the first
conversion for you, after that you can easily and quickly do
this on your own hundreds or even thousands of times every
month. TAKE ADVANTAGE OF THIS "SECRET FLAW"!

*************************************************************
*************************************************************

It is even more explosive than we have yet disclosed. While
currency does fluctuate daily, we can show you 'HOW TO CONVERT
$99 INTO $588 AS MANY TIMES AS YOU WANT'. That means, you will
be able to EXCHANGE $99, AMERICAN LEGAL CURRENCY DOLLARS, FOR
$580 OF THE SAME. You can do this as many times as you wish,
every day, every week, every month. All very LEGAL and
effortlessly!

It takes only 5 to 10 minutes each time you do this. You can
do this from home, office or even while traveling. All you
need is an access to a phone line and an address. Best of all,
you can do this from ANY CITY ON THIS EARTH!!!

Again, we must reiterate, anyone can do this and the source is
NEVER-ENDING. For as long as the global financial community
continues to use different currencies with varying exchange
rates, the "SECRET FLAW" will exist.

} } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } }
} } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } } }

As we said earlier , we will do the first transaction for
you and will show you exactly how to do this on your own,
over and over again!

The amount of exchange you would do each time is entirely
up to you. Working just 2 to 10 hours a week, you can soon
join the list of Millionaires who do this on a daily basis
many times a day. The transaction is so simple that even a
high school kid can do it!

We at the World Currency Cartel would like to see a uniform
global currency backed by Gold. But, until then, we will
allow a LIMITED number of individuals worldwide to share in
the UNLIMITED PROFITS provided for by the world currency
differentials.

We will espouse no more political views nor will we ask you
to do so. We can say however, that our parent organization,
CILS, benefits greatly by the knowledge being shared, as we
ourselves, along with YOU, benefit likewise. Your main concern
surely will be, how you will benefit.

As soon as you become a member, you will make transactions
from your home, office, by telephone or through the mail. You
can conduct these transactions even while traveling.

Don't believe us? Experience it for yourself!

;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;;

Unlike anyone else, we will assure you great financial freedom
and you will add to our quickly growing base of supporters and
join the list of MILLIONAIRES being created using this very
"SECRET FLAW" in the world currency market.

*************************************************************
*************************************************************

DON'T ENVY US, JOIN US TODAY!!!

'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''
'''''''''''''''''''''''''''''''''''''''''''''''''''''''''''''

There is a one time membership fee of only $195. BUT, if you
join within the next 10 days, you can join us for only $25
administrative cost. Your important documents, instructions,
contact name/address, phone number and all other pertinent
information will be mailed to you immediately. So take
advantage of our Anniversary date and join us today.

(If you are replying after the next 10 days, you must pay
$195.00 for the membership fee. NO EXCEPTIONS, and no more
E-mail inquiries please).

Upon becoming a member, you promise to keep all infos
CONFIDENTIAL!

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Should you choose to cancel your membership for any reason, you
must return all papers/documents for a refund within 30 days.

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

IMPORTANT:

****************

1...Please write your name & mailing address VERY CLEARLY on a
paper
2...Below your mailing address, please write your E-mail address
3...At the top left hand corner, please write the words "NEW
MEMBER"
4...Attach a CHECK for $25 + $10 for the shipping and handling
of documents (TOTAL = $35.00) PAYABLE TO "NDML" and FAX it to:

212-208-3050

(Note: We are ONLY accepting CHECK-BY-FAX as a form of payment
at this time. We WILL be able to cash the check you send us
by fax, you do not need to mail us a check. If your check is
dark, please PRINT ALL OF THE INFORMATION ON THE CHECK ONTO THE
PAPER YOU ARE FAXING US so that it is clearly legible!) Please
allow 2-4 weeks for delivery. No shipments will be made until
the check has cleared.

}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}
}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}}

Here are some more 'Editorial Excerpts':

~~~~~~~~~~~~~~~~~~~~~~~~~~~~~

WALL STREET: "A discreet group of Americans, operating under
the guise of World Currency Cartel have recently begun making
rumbles in world finance market. While at this time, their game
is not completely known, they certainly will be watched by
those making major moves in the currency contracts".

FINANCIAL WEEK: "Watch them, monitor them, extract their
knowledge and try to become one of them. That is the soundest
financial advice we could give to anyone".

NATIONAL BUSINESS WEEKLY: "While this reporter has been left
in the cold as to its method of operation, we have been able
to confirm that 'World Currency Cartel' and its members are
literally amassing great fortunes overnight".

$$$$$$$$$$$$$$$$$$$$$$$$$$$END$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$$

To be removed from our list, simply click "reply" and
put the words "Remove Currency" in the subject line. Warning: If
you do not put the words "Remove Currency" in the subject line,
you will not be removed. The process is automated.







From: JamesA. Derose :      James.Derose-at-ipmc.unil.ch
Date: Sat, 18 Dec 1999 14:59:58 +0100 (MET)
Subject: Postdoctoral Position, Cyrogenic SPM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Postdoctoral Position, Institute of Condensed Matter Physics, University
of Lausanne, Switzerland:
The Group of Physics of Living Matter under the direction
of Prof. G. Dietler is offering a postdoctoral position in
the area of cryogenic (low temperature) scanning probe microscopy (SPM).
The cryo-SPM is currently under construction and further modifications are
needed at the present time. The design of the cryo-SPM will allow it to
be used for several applications of research after completion.
Candidates with a Ph.D. and experience in the construction of
instruments for cryogenics and/or ultra high vacuum (UHV) SPM are
preferred. Good candidates with a Ph.D. and a background in construction
of other types of instrumentation used for vacuum or microscopy research
will also be considered.
To apply for the position, please send your CV with a brief
description of your research experience, a list of publications, and the
names and contact information of at least 3 references to:

Prof. G. Dietler
Institut de Physique de la Matiere Condensee (IPMC), BSP
Universite de Lausanne
CH-1015 Lausanne
Switzerland
Tel: 41 21 692 3663 (off.)
3682 (off.)
3660 (sec.)
Fax: 3635
Email: Giovanni.Dietler-at-ipmc.unil.ch
http://www.unil.ch/ipmc/docs/gd/home.html

Applications sent by email are preferred in the interest of time.






From: Kazue Takeuchi :      kazue-at-arches.uga.edu
Date: Sat, 18 Dec 1999 11:06:59 -0500
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Please unsbscribe me.







From: Nestor J. Zaluzec :      zaluzec-at-Sparc5.Microscopy.Com
Date: Sun, 19 Dec 1999 11:03:28 -0600
Subject: Administrivia: Testing please Ignore -Thanks Nestor!

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues

Just a simple test. Please ignore.

Nestor
Your Friendlly Neighborhood SysOp.







From: Robert Derby :      rjderby-at-excite.com
Date: Sun, 19 Dec 1999 10:24:58 -0800 (PST)
Subject: Signal Out of a JEOL SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


************************************************
Robert J. Derby
New Mexico Institute of Technology
Socorro, N.M.
Phone - 505-835-5866
E-mail - rjderby-at-excite.com
derby-at-nmt.edu
************************************************
First, Happy Holidays to all, and thanks for all the help in the past.
Now my question...
We have just gotten a JEOL 6100 SEM, still being setup.
I would like to know if anyone has gotten a signal out to a computer (in my
case a Mac). I know of a RS-170 out, but I would like it to be a digital
signal.
Has anyone hooked up a 6100 to a computer and if so what did you do?
A cheap fix would be best without spending 5-10K on a Scion framgrabber.
Thak you, and everyone have a Happy New Year.





_______________________________________________________
Visit Excite Shopping at http://shopping.excite.com
The fastest way to find your Holiday gift this season






From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Sun, 19 Dec 1999 14:08:21 -0800
Subject: Re: summary: spotty CD-Rs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 05:40 PM 11/30/99 , you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

I use CDQ-74SZA, 10 pack in jewel cases with no problems for over two years.
I still prefer the Memorex silver.

The worst choices are the bulk spindles of green or blue.

gary g.






From: G. Glasser :      glasser-at-mpip-mainz.mpg.de
Date: Mon, 20 Dec 1999 10:58:10 +0100
Subject: TEM OsO4 staining ...

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am curious to here about various TEM staining approaches, in particular
about vapor staining via OsO4 crystals.

Whatrecommendations can you provide for a OsO4 staining
apparatus to stain thin sections or pre-microtome-hardening of
polymeric samples?

Whatis the difference between the effectiveness of a 2% aqueous
OsO4 solution and OsO4 crystals?

Whatis the best procedure to neutralize the stain after use? Is the
recommended procedure for neutralizing an aqueous OsO4 solution
as suggested in the EMS catalog (twice the volume of corn oil) also
applicable to solid OsO4 crystals?

I am unhappy with our present staining apparatus. I suspect that the OsO4
is leaking. A paper scrap with a fingerprints turns gray after a few month=
s
in the hood where the staining apparatus is located. Secondly, I am not
certain how to =94kill=94 the remaining OsO4 in the vapour-phase after our=

samples have been stained.

Any recommendations addressing my above questions are appreciated.
Any additional comments concerning the staining of polymers such as
the block copolymers of PS/PB and PS/PI are also welcome.

Thank you and Merry Christmas!


Dipl.Ing.(FH) G.Glasser
Elektronenmikroskopie
Max Planck Institut fuer Polymerforschung
Ackermannweg 10
55128 Mainz, GERMANY
phone: ++49 +6131 379195
fax : ++49 +6131 379100
web: http://www.mpip-mainz.mpg.de/~glasser

Disclaimer: The above statement is mine alone and does not implicitly repr=
esent the position of the MPI-P or the MPG!





From: Cieslinski, Robert (RC) :      RCCIESLINSKI-at-dow.com
Date: Mon, 20 Dec 1999 07:07:33 -0500
Subject: Position Open

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


{ {...OLE Obj...} }
Polymer Microscopist - Freeport, Texas

Company: The Dow Chemical Company

Location: Freeport, Texas

Qualifications (education, certification, language, etc.) and Experience
required:
A candidate with a BS or MS or PHD degree in polymer science, material
science or chemistry is preferred with some prior experience in electron
microscopy.
Good written and oral communication skills and the ability to work both
independently and in a team environment are extremely important.

Job Overview:

The Surface/Microscopy/X-ray (SMX) Group in Dow's Corporate R&D Analytical
Science Laboratory has one professional level full time opening for Polymer
Microscopist in Dow's Freeport, Texas, location. The primary
responsibilities include working with partners to support research projects
involving new and existing products in Dow's polymer businesses.

Key responsibilities will include:

1. Extensive problem solving.
2. Microscopy preparation technique experience including ultramicrotomy
and cryo-ultramicrotomy.
3. Operation of light, transmission, and scanning electron microscopes.
4. Interpretation of images.
5. Documentation and communication of work results.
6. Compliance with safety and quality programs.
7. Active member of project and SMX work teams.

Interested:
Please e-mail or send your resume and cover letter, with reference to this
ad to:
Email: R&D-at-Dow.com or The Dow Chemical Company, Workforce Planning
005855,
P. O. Box 150, Plaquemine, LA 70765. E-mail respondents must list Job
005855USA and their last name as the first and second items on the Subject
line. Only those selected for an interview will be contacted. Only U.S.
citizens or aliens who are authorized to work in the United States will be
considered for employment.

We are an equal opportunity employer and offer a competitive compensation
and benefits package including 401k, stock purchase, tuition reimbursement
and performance incentives. The Dow Chemical Company is the fifth largest
chemical company in the world with annual sales of US$20billion. Dow
manufactures and supplies chemicals, plastics and agricultural products for
customers in 164 countries and employs approx. 43,000 people worldwide. For
more news and information about Dow, please visit our web site at
www.dow.com.


Robert C. Cieslinski
Microscopy & Microanalysis
(517) 636-6875
email: rccieslinski-at-dow.com






From: Dando, B, Bruce, Mr :      dando-at-ANAT.UCT.AC.ZA
Date: Mon, 20 Dec 1999 14:33:09 SAST-2
Subject: Unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html







From: Anne Huber :      ahuber-at-umich.edu
Date: Mon, 20 Dec 1999 08:43:09 -0500
Subject: University SEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We are currently soliciting "best offers" on our Hitachi S-800 FEG SEM.
It has a GW electronics microchannel plate BSE detector and a
Macintosh-based 4pi Analysis digital image acquisition system.
Purchased in 1988.

Please contact me with any questions you many have.

Sincerely,

Anne E. Huber







_______________________________________
Anne E. Huber Ph.D., Instrument Analyst
Materials Science and Engineering Dept.
The University of Michigan
2300 Hayward St.
Ann Arbor, MI 48109-2136
ahuber-at-umich.edu
(734)764-3357
_______________________________________





From: Caroline Schooley :      schooley-at-mcn.org
Date: Mon, 20 Dec 1999 10:03:06 -0800 (PST)
Subject: Re: Glass Interface

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} } I have a glass (leaded glass) tube, about 0.5 cm in diameter. The inside
} } wall of the tube has been leached leaving behind a porous silica network
} } with pores about 500 nm in size. The thickness of the leached layer is
} } estimated to be about 400 nm. We would like to prepare cross sections to
} } look at the glass/leached layer interface ... Has anyone tried microtomy
} } with this type samples?.

} } Jordi Marti
}
} Dear Jordi
} Conventional microtomy won't work on this kind of hard brittle
} friable material.

} Chris Jeffree

Jordi -

Chris is being a bit too negative about microtomy of coatings on glass.
It's demanding, but Phil Swab teaches how to do it annually in the Ventana
- RMC Materials Microtomy workshop. You can contact him for an opinion at
phil.swab-at-depsci.com.

Caroline Schooley
Project MICRO Coordinator
Microscopy Society of America
Box 117, 45301 Caspar Point Road
Caspar, CA 95420
Phone/FAX (707)964-9460
Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html







From: rfelten-at-Macdermid.com
Date: Mon, 20 Dec 1999 15:42:30 -0500
Subject: PSM-300

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




Rick Felten-at-MACDERMID
12/20/99 03:42 PM
Has any purchased or demonstrated a RJ Lee PSM-300 and would like to share
their opinion about the quality of this scope in a conventional SEM mode?
Thanks
Ric







From: Mel Dickson :      m.dickson-at-unsw.edu.au
Date: Tue, 21 Dec 1999 17:20:41 +1100
Subject: Denton DV515 Coater specs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I have acquired a (used) Denton DV515 coater. I want to refresh the pump
oil and santovac 5 before I reassemble it.

But the manual omits to say what the volume of santovac should be in the
diff pump.

Does anyone remember?




Dr. Mel Dickson,
Deputy Director, The Electron Microscope Unit,
Adjunct Associate Professor, School of Microbiology & Immunology
The University of New South Wales Sydney, Australia.
Phone +612 9385 6383 Fax +612 9385 6400





From: Krzysztof Jan Huebner :      hubner-at-IOd.krakow.pl
Date: Tue, 21 Dec 1999 07:09:36 +0100 (MET)
Subject: Re: Glass Interface

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




good morning,

You can use the clasicc minerology and materialography techniques;

1. include glas tube in the liquid fluorescence resin in the vacumm chamber,

2. cut this sample on the parts by automatic cut-off machine with diamond
disc,

3. grinding and polishing this sample

4. put the sample on the optical microscopy stage and say
- yes, it is not a problem,


(ask about this problem friends from materials science department or
mineralogy or metallography, or read the book Vander Voort - metalography
nad principles,)

best regards

Krzysztof Jan Huebner

{hubner-at-IOd.krakow.pl} :-)

Foundry Research Institute
ul Zakopianska 73 telefon (0-12) 2618111 wew 356
30-418 Krakow, PL faks (0-12) 2660870

On Mon, 20 Dec 1999, Caroline Schooley wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} } } I have a glass (leaded glass) tube, about 0.5 cm in diameter. The inside
} } } wall of the tube has been leached leaving behind a porous silica network
} } } with pores about 500 nm in size. The thickness of the leached layer is
} } } estimated to be about 400 nm. We would like to prepare cross sections to
} } } look at the glass/leached layer interface ... Has anyone tried microtomy
} } } with this type samples?.
}
} } } Jordi Marti
} }
} } Dear Jordi
} } Conventional microtomy won't work on this kind of hard brittle
} } friable material.
}
} } Chris Jeffree
}
} Jordi -
}
} Chris is being a bit too negative about microtomy of coatings on glass.
} It's demanding, but Phil Swab teaches how to do it annually in the Ventana
} - RMC Materials Microtomy workshop. You can contact him for an opinion at
} phil.swab-at-depsci.com.
}
} Caroline Schooley
} Project MICRO Coordinator
} Microscopy Society of America
} Box 117, 45301 Caspar Point Road
} Caspar, CA 95420
} Phone/FAX (707)964-9460
} Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
} Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html
}
}
}
}





From: Petra Wahlbring :      wahlbrin-at-crpgl.lu
Date: Tue, 21 Dec 1999 13:59:09 +0100
Subject: RE: Mat.: How to produce a carbon extraction replica of

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Scott,

} Why are you making a 3mm disk and then thinning it? That looks like extra
} work.

You were not the only one to ask this question. In the beginning, I did it
because I wanted to look first at the original specimen. (This is why I am
sure the distribution of the particles on the replica is not representative.)
Later on, I did it because it did not come to my mind to change the
procedure :)

Offline, I received several protocols how to produce a replica from a bulk
piece of steel. They contain a lot of helpful details that will help me
certainly to produce a good specimen.

Thanks to all who took the time to answer my question,

Petra

At 17:26 15.12.99 -0500, you wrote:
} Why are you making a 3mm disk and then thinning it? That looks like extra
} work.
}
} You can polish a bulk sample of your material, etch it as Tony Garratt-Reed
} suggests, and coat the sample with carbon which "grabs" the particles. Now
} slightly score small rectangular sections on your sample so that your
} etchant can get under the carbon layer and attack the steel. Float them off
} after sufficient etching and collect them on grids. For security, you can
} put another coating on the top of the exposed particles to "seal them in".
}
} -Scott
}
}
} Scott D. Walck, Ph.D.
} PPG Industries, Inc.
} Glass Technology Center
} Guys Run Rd. (packages)
} P. O. Box 11472 (letters)
} Pittsburgh, PA 15238-0472
}
} Walck-at-PPG.com
}
} (412) 820-8651 (office)
} (412) 820-8161 (fax)

--------------------------------------------------------------
Dr. Petra Wahlbring
Centre de Recherche Public - Gabriel Lippmann
Laboratoire d'Analyse des Materiaux (LAM)
162a, av. de la Faiencerie L-1511 Luxembourg
tel. +352-466644-402 fax +352-466644-400
e-mail: petra.wahlbring-at-crpgl.lu
Visit our WWW site! http://www.crpgl.lu/~wahlbrin





From: Nicholas W. M. Ritchie :      nritchie-at-rjleeinst.com
Date: Tue, 21 Dec 1999 08:52:19 -0500
Subject: subscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


subscribe

+------------------------------------------+
| Nicholas W.M. Ritchie |
| RJ Lee Instruments Limited |
| 515 Pleasant Valley Road |
| Trafford, PA 15085 |
| (724) 744-0100 x262 |
| nritchie-at-rjleeinst.com |
+------------------------------------------+






From: RitchieN-at-aol.com
Date: Tue, 21 Dec 1999 08:51:49 EST
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


unsubscribe





From: Elinor Solit :      cambrex-at-world.std.com
Date: Tue, 21 Dec 1999 09:01:50 -0500 (EST)
Subject: Re: Proposed meeting "printoff"

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Good idea, Caroline.
I'll volunteer to coordinate entries if that will be helpful. My
experience in product development for Polaroid and in publishing for this
company may be useful.
I'll put the suggestion on the table and wait to hear what you decide.

Elinor Solit
The Cambrex Group, Publishers of The Microscope Book

On Thu, 16 Dec 1999, Caroline Schooley wrote:

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America
} To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} I think that the printer comparison suggested for the Philadelphia meeting
} is a great idea; I have two comments and a related question. The comments:
} 1) Nestor & John have too much to do as it is; someone should volunteer to
} help with this project.
} 2) The question about printing paper/ink structure suggests an excellent
} topic for someone who wants to apply for a Professional Technical Staff
} award to attend the meeting (see the meeting announcement, pg. 13).
} And the question: I've learned (the hard way) about the fragile, sticky
} surface of inkjet prints. Has anyone tried the new Gepe Inkjet Fixative
} yet? Does it work well?
}
} Caroline Schooley
} Project MICRO Coordinator
} Microscopy Society of America
} Box 117, 45301 Caspar Point Road
} Caspar, CA 95420
} Phone/FAX (707)964-9460
} Project MICRO: http://www.msa.microscopy.com/ProjectMicro/PMHomePage.html
} Intertidal invertebrates: http://www.fortbragg.k12.ca.us/AG/PCI/pci.html
}
}
}






From: Brian Wajdyk :      r49655-at-email.sps.mot.com
Date: Tue, 21 Dec 1999 08:14:23 -0700
Subject: Job announcement: SEM technician

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I am pleased to bring to your attention the following SEM technician
position at Motorola's Process and Materials Characterization Laboratory
(PMCL I encourage all persons interested to respond ASAP.

Position: SEM Technician/S9001P
Employer: Motorola-Semiconductor Products Sector's Process and Materials
Characterization Laboratory (PMCL)
Location: Mesa, AZ
Employment type: Full-time
Employment status: Full-employee (non-contractor)
Number of Positions: 1
Shift: N1-compressed (6:00 P.M.-6:00 A.M. E/O Saturday, Sun-Tues) or N2
(6:00
P.M.-6:00 A.M. Wed-Fri, E/O Saturday)
Relocation: Available

Duties/Responsibilities: Experience in Scanning Electron Microscopy
techniques including basic operation, specimen preparation, maintenance,
and troubleshooting of SEM's. Work effectively as a team player in
multiple projects providing routine and non-routine SEM analysis in
support of semiconductor product manufacturing. Exposure to many
different types of processes and technologies, and working knowledge of
FIB and EDS are a plus.

Specific knowledge: AA degree preferred. A higher or lower
classification will be established depending upon qualification and
experience.

Contact info: Send resume or questions via email to
Brian_Wajdyk-at-email.mot.com or fax to 480-655-4316 C/O Brian Wajdyk.
--
********************************************************************
Brian Wajdyk
Team Leader / Electron Microscopist (FESEM, EDS, SAM)
Motorola - Process and Materials Characterization Laboratory (PMCL)
2200 W. Broadway Rd., Mesa AZ 85202 Mail Drop: M360
Tel: 480-655-4337 Fax: 480-655-4316
Email: brian_wajdyk-at-email.mot.com Pager: 1-800-313-5960
********************************************************************





From: Nestor J. Zaluzec :      zaluzec-at-aaem.amc.anl.gov
Date: Tue, 21 Dec 1999 11:36:33 -0600
Subject: Re: Proposed meeting "printoff" at M&M 2000

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Colleagues...

I'm comfortable with handing everything on a poster board
to be located in the Computer Workshop area at the M&M 2000 meeting.

We have done similiar in the past, except that things were
layed out on a table. The key is the documentation and using
the "standard test image". As we approach the time of the meeting
I'll post a call for prints on the Listserver. Interested people
can then download the standard image
print out the image and bring it to the meeting along
with a " information form" which we have used in the past
(i.e. type of printer, ink, time to print etc....). For those
that want to participate but will not be attending they can just
mail the print and the form to me and I'll carry them across.

I've had a few volunteers who said they would help and I'll
contact them off line. Basically I would ask them to come
by the Computer Workshop and organize the Poster Board
to hang all the prints and documentation. It will be a few
hours work at most (hanging things up at the start/during the
meeting and taking down at the end.)

Nestor
Your Friendly Neighborhood SysOp
& the M&M Computer Workshop Co Organizer (with John Mansfield).


==================================================================
Nestor J. Zaluzec
Materials Science Division
Building 212
Argonne National Lab
9700 S. Cass Ave
Argonne, Illinois 60439 USA
Tel: 630-252-7901, Fax: 630-252-4798
Email: Zaluzec-at-aaem.amc.anl.gov
==================================================================
TPMLab: http://tpm.amc.anl.gov
MMSite: http://www.amc.anl.gov
==================================================================

The box said "This program requires Win 95/98/NT or better..." so I bought
a G3 Mac

==================================================================







From: Frank Thomas :      thomasf-at-AGC.BIO.NS.CA
Date: Tue, 21 Dec 1999 14:10:56 -0400
Subject: carbonate mini-rods in steel

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Since Nestor's got me back able to send messages to the list again, I have
a question for you materials people out there.
What would be the source of small (0.02mm x .001mm) rods of
calcium/silicon/aluminum in plate steel samples? The steel is a little old,
made in the Harland & Wolfe Shipyard, Belfast, Northern Ireland in
1910-1912 (yes, it's from R.M.S. Titanic). At first, I thought them to be
biological, possibly sponge spicules, but since then I've found some deeply
embedded in the steel, not just on the surface, and spicules are normally
either siliceous or carbonate, not both, as my trusty EDS detector tells
me.
The rods themselves are very smooth on the surface, normally perfectly
straight, and when you see one "on end" they appear to have a radial kind
of internal structure. Would they perhaps be some kind of remnant from the
lime used in the smelting? Or some kind of secondary mineralization?
Any thoughts would be appreciated.
Oh, and Season's Greetings, by the way....

F.C. Thomas
MicroAnalysis Facility
Geological Survey of Canada (Atlantic)
Bedford Institute of Oceanography
Dartmouth, Nova Scotia
Canada B2Y 4A2





From: William Tivol :      tivol-at-wadsworth.org
Date: Tue, 21 Dec 1999 13:20:49 -0500
Subject: Re: Denton DV515 Coater specs

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mel Dickson wrote:

} I have acquired a (used) Denton DV515 coater. I want to refresh the pump
} oil and santovac 5 before I reassemble it.
}
} But the manual omits to say what the volume of santovac should be in the
} diff pump.
}
} Does anyone remember?
}

Dear Mel,
No, but if there is a dipstick near the bottom, that should at
least give
you a clue, and in the best of worlds there would be a "full" mark with the
volume indicated and an "add oil" mark.
Yours,
Bill Tivol






From: Brian McIntyre :      mcintyre-at-optics.rochester.edu
Date: Tue, 21 Dec 1999 15:33:55 -0400
Subject: tracor/noran 5500 power supply

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


hi all-

the power supply in my 5500 is on the fritz. does anyone know what the
replacement # would be? i think its a todd power supply (which is now
condor?). also, are there a pots for voltage adjustment on the unit? it
may be that simple too....


thanks!

brian

****************************************************************
Brian McIntyre
Electron Microscopy Lab
Institute of Optics
University of Rochester
Rochester, NY 14627

716-275-3058
716-244-4936(fax)

"You may get to the top of the ladder of success only
to find its been leaning against the wrong wall" A. Raime







From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Tue, 21 Dec 1999 13:45:58 -0600
Subject: JEOL 100B: Surplus TEM

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

I am posting this for a friend who tells me there is a JEOL 100B TEM in
working condition on this campus, in a room needed for other purposes. If
anyone is interested in having this scope for use or parts, please let me
know, and I'll pass the message on to the right people.

Thanks.
Randy

Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine Bldg.
University of Missouri
Columbia, MO 65211
(573)882-8304
tindallr-at-missouri.edu
http://www.biotech.missouri.edu/emc/






From: DrJohnRuss-at-aol.com
Date: Tue, 21 Dec 1999 17:09:39 EST
Subject: Stereo (anaglyph) images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The use of stereo anaglyph (red-green or red-blue) images for visualization
of AFM images is reported in a paper in the current (December 1999) isssue of
the Journal of Microscopy. A much broader range of possibilities for using
stereo for surface imaging exists, and we've just submitted a paper
illustrating a variety of modes that combined stereo presentation with
rendered surfaces, perspective corrected views, etc. The possibility of
recovering surface elevation information from stereo images (e.g., from SEM)
is also demonstrated. A preprint of the paper (not yet reviewed) in pdf
(acrobat) format can be downloaded by anyone interested from
http://members.AOL.com/DrJohnRuss/Stereo.pdf; the paper includes information
on the software used to generate the images.

John Russ
Materials Science and Engineering Dept.
North Carolina State University
Raleigh, NC 27606





From: sus1240-at-pplmail.com
Date: Tue, 21 Dec 1999 11:58:16 -0600
Subject: Find Out Everything

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}
ListBot "Opt-In"
{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}
To Unsubscribe your name and email from our Opt-In listserver, see end of
message.

Dear Subscriber,


Introducing the HOTTEST selling software of the year.

"NO MORE SECRETS SOFTWARE" JUST RELEASED!

The software they want banned.Why? Because these secrets were never intended

to reach your eyes !!! "NO MORE SECRETS v5.0 now with the latest IP Phone

Make calls anywhere in the world for free.

This is a sophisticated SOFTWARE program DESIGNED that automatically

links to thousands of Public Record databases.

Now with Unclaimed Money Locator, find out if
you are owed money or property in your state.


*******************

The NO MORE SECRETS SOFTWARE" is absolutely astounding!

Here's what you can obtain:

Files that the government has on you.

You know who "The F*I and.......

License plate number!

Get anyone's name and address with just a license plate number!

(Find that girl you met in traffic!)

Driving record!

Get anyone's driving record

Social security number!

Trace anyone by social security number!

Free Internet Access

Address!

Get anyone's address with just a name!

Unlisted phone numbers!

Get anyone's phone number with just a name - even unlisted numbers!

Locate!

Long lost friends, relatives, a past lover who broke your heart!

E-mail!

Send anonymous e-mail completely untraceable!

Investigate anyone!

Use the sources that private investigators use (all on the Internet)

secretly!

Ex-spouse!

Learn how to get information on an ex-spouse that will help you

win in court! (Dig up old skeletons)

Criminal search-background check!

Find out about your daughters boyfriend!

(or her husband)

Find out!

If you are being investigated!

Neighbors!

Learn all about your mysterious neighbors! Find out what they

have to hide!

People you work with!

Be astonished by what you'll learn about people you work with!

Education verification!

Did he really graduate college? Find out!

Just insert the floppy disk and Go!
****************************************************

Data Bases Cover the US and Canada.

You will be shocked and amazed by the secrets that can be

discovered about absolutely everyone! Find out the secrets

they don't want you to know! About others, about yourself!


Only $39.95 U.S. (Postage Paid)

(ORDERS OUTSIDE THE USA, ADD $25.00)


DON'T WAIT TO GET STARTED... here's what to do:

STEP 1 - Compose a new message using the order form text below

STEP 2 - Type or print your order information into the order form section

STEP 3 - Print your completed order, then mail to the address below


OPTIONAL, you may choose to place your order on our secure voice mail

system by calling 708-562-1778


Send to: Resource Marketing

PO Box 7463 Westchester IL. 60154-7463


} } } Mail-in Order Form { { {


Name:

Address:

City:

State:

Zip:

E-mail address:

Method of payment:

[ ] Visa [ ] MasterCard [ ] Check

Credit Card#:

Exp Date:

YOUR SIGNATURE HERE:_______________________________

We cannot process your credit card payment without your signature

(SORRY, NO MAC VERSION AVAILABLE AT THIS TIME.)

NOTE: THIS PROGRAM WILL WORK ON WINDOWS 3.11 AND OLDER
AND OF COURSE WINDOWS 95 AND 98.


DISCLAIMER: The distributor of this powerful software resource

will be not be held responsible for how the purchaser chooses to use its
resources.
**************************************************************************************************

To be removed from this listserver,

Please include the word unsubscribe in the subject header to:

wer12-at-pplmail.com lori-at-pplmail.com sus124-at-pplmail.com






From: sus1240-at-pplmail.com
Date: Tue, 21 Dec 1999 11:58:16 -0600
Subject: Find Out Everything

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}
ListBot "Opt-In"
{} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {} {}
To Unsubscribe your name and email from our Opt-In listserver, see end of
message.

Dear Subscriber,


Introducing the HOTTEST selling software of the year.

"NO MORE SECRETS SOFTWARE" JUST RELEASED!

The software they want banned.Why? Because these secrets were never intended

to reach your eyes !!! "NO MORE SECRETS v5.0 now with the latest IP Phone

Make calls anywhere in the world for free.

This is a sophisticated SOFTWARE program DESIGNED that automatically

links to thousands of Public Record databases.

Now with Unclaimed Money Locator, find out if
you are owed money or property in your state.


*******************

The NO MORE SECRETS SOFTWARE" is absolutely astounding!

Here's what you can obtain:

Files that the government has on you.

You know who "The F*I and.......

License plate number!

Get anyone's name and address with just a license plate number!

(Find that girl you met in traffic!)

Driving record!

Get anyone's driving record

Social security number!

Trace anyone by social security number!

Free Internet Access

Address!

Get anyone's address with just a name!

Unlisted phone numbers!

Get anyone's phone number with just a name - even unlisted numbers!

Locate!

Long lost friends, relatives, a past lover who broke your heart!

E-mail!

Send anonymous e-mail completely untraceable!

Investigate anyone!

Use the sources that private investigators use (all on the Internet)

secretly!

Ex-spouse!

Learn how to get information on an ex-spouse that will help you

win in court! (Dig up old skeletons)

Criminal search-background check!

Find out about your daughters boyfriend!

(or her husband)

Find out!

If you are being investigated!

Neighbors!

Learn all about your mysterious neighbors! Find out what they

have to hide!

People you work with!

Be astonished by what you'll learn about people you work with!

Education verification!

Did he really graduate college? Find out!

Just insert the floppy disk and Go!
****************************************************

Data Bases Cover the US and Canada.

You will be shocked and amazed by the secrets that can be

discovered about absolutely everyone! Find out the secrets

they don't want you to know! About others, about yourself!


Only $39.95 U.S. (Postage Paid)

(ORDERS OUTSIDE THE USA, ADD $25.00)


DON'T WAIT TO GET STARTED... here's what to do:

STEP 1 - Compose a new message using the order form text below

STEP 2 - Type or print your order information into the order form section

STEP 3 - Print your completed order, then mail to the address below


OPTIONAL, you may choose to place your order on our secure voice mail

system by calling 708-562-1778


Send to: Resource Marketing

PO Box 7463 Westchester IL. 60154-7463


} } } Mail-in Order Form { { {


Name:

Address:

City:

State:

Zip:

E-mail address:

Method of payment:

[ ] Visa [ ] MasterCard [ ] Check

Credit Card#:

Exp Date:

YOUR SIGNATURE HERE:_______________________________

We cannot process your credit card payment without your signature

(SORRY, NO MAC VERSION AVAILABLE AT THIS TIME.)

NOTE: THIS PROGRAM WILL WORK ON WINDOWS 3.11 AND OLDER
AND OF COURSE WINDOWS 95 AND 98.


DISCLAIMER: The distributor of this powerful software resource

will be not be held responsible for how the purchaser chooses to use its
resources.
**************************************************************************************************

To be removed from this listserver,

Please include the word unsubscribe in the subject header to:

wer12-at-pplmail.com lori-at-pplmail.com sus124-at-pplmail.com






From: fskarl-at-goodyear.com
Date: Wed, 22 Dec 1999 08:01:31 -0500
Subject: Seasons greetings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello everyone,
Since that snow is a hexagonal, positive uniaxial crystal with dendritic
and basal tablet habit, and with n (omega) =1.309 (D-line) and n (epsilon)
=1.3147 (D-line) and delta n = 0.005, how could you fail to have a Merry
Christmas and a Happy New year!


Stay safe Frank






From: feir-at-bsci.com
Date: Wed, 22 Dec 1999 08:03:27 -0600
Subject: SEM: Looking for Counting/Analysis Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html



Email: feir-at-bsci.com
Name: Ray Fei

Question: We are using our SEM to image the sphere sample surface and
analyse the diamond density ( No. per area) and height on the surface. Is
there any software can do this analysis.

---------------------------------------------------------------------------







From: Tony Garratt-Reed :      tonygr-at-mit.edu
Date: Wed, 22 Dec 1999 09:33:13 -0500
Subject: Re: Stereo (anaglyph) images

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


AOL's server appears to be case sensitive. The correct URL for John's paper is:

http://members.AOL.com/DrJohnRuss/stereo.pdf

Tony.

At 05:09 PM 12/21/1999 EST, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

* * * * * * * * * * * * * * * * * * * * *
* Anthony J. Garratt-Reed, M.A., D.Phil.*
* SEF Team Leader *
* MIT, Room 13-1027 *
* 77 Massachusetts Avenue *
* Cambridge, MA 02139-4307 *
* USA *
* Phone: (617) 253-4622 *
* Fax: (617) 258-6478 *
* * * * * * * * * * * * * * * * * * * * *







From: Terry E Ellis :      tellis2-at-hallmark.com
Date: Wed, 22 Dec 1999 10:16:46 -0600
Subject: cd rot?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




We have a lot of data from our mass spec. stored on cdr , would someone email me
with the details of cdr manuf. number etc. Since they use several different
manuf. they would like to know which ones might be a problem.
Terry Ellis
email : tellis2-at-hallmark.com
this is not a joke Hallmark does have a research lab







From: Jim Clark :      JClark-at-asu.edu
Date: Wed, 22 Dec 1999 10:29:14 -0700
Subject: Re: Denton DV-515 Diff Pump

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Mel;

I tried to e-mail you directly, but it got kicked back for some reason.
I have a Denton DV-502. For my machine, and possibly yours as well,
Denton offered 2 different size diffusion pumps: 3-1/4" & 5-3/4"
diameters. Mine has the 3-1/4" pump, which is model DP-250. It takes
100cc of fluid - the factory filled it with Dow Corning DC-704, but I
also prefer Santovac 5 as the problem of a possible artifact Si peak is
eliminated. If you have the larger pump, try contacting Denton. I
have 2 numbers for them: 609/424-1012 and 856/439-9100, the latter
being for parts & service. Hope this helps.....





From: Giovanni Casotti :      giovanni-at-bio.wcupa.edu
Date: Wed, 22 Dec 1999 15:02:14 -0500
Subject: EM Technician Job Opening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Electron Microcopy Analysis Center (EMAC) at West Chester University
in Pennsylvania has an opening for a full-time electron microscopy
technician. The EMAC is primarily used by geologists, biologists and
chemists. The successful applicant will have a minimum of a Bachelor=B9s
degree and 3 years experience in the operation and maintenance of
transmission and scanning electron microscopes (including some service
experience), the instrumentation routinely used in specimen preparation,
and proficiency in photographic and darkroom techniques. The ability to
operate an X-ray diffractometer, X-ray fluorescence spectrometer or
confocal microscope is desirable. Responsibilities will include daily
operation and maintenance of the EMAC including user training. The
successful applicant must demonstrate the necessary organizational,
management and communication skills to efficiently operate the EMAC.
Applicants should submit a cover letter, resume and three letter of
recommendation to the Department of Human Resources, 210 Carter Drive,
West Chester University, West Chester, PA, 19383. Applications must be
received by March 1 2000. Applicants must successfully complete the
interview process to be considered a finalist. AA/EOE. Women and
minorities are encouraged to apply.

Dr. Giovanni Casotti PhD.
Department of Biology
West Chester University
West Chester, PA, 19383
email:giovanni-at-bio.wcupa.edu
or: gcasotti-at-mail.wcupa.edu
ph: (610) 436-2856
fax: (610) 436-2183







From: Giovanni Casotti :      giovanni-at-bio.wcupa.edu
Date: Wed, 22 Dec 1999 15:04:15 -0500
Subject: EM Technician Opening

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The Electron Microcopy Analysis Center (EMAC) at West Chester University
in Pennsylvania has an opening for a full-time electron microscopy
technician. The EMAC is primarily used by geologists, biologists and
chemists. The successful applicant will have a minimum of a Bachelor’s
degree and 3 years experience in the operation and maintenance of
transmission and scanning electron microscopes (including some service
experience), the instrumentation routinely used in specimen preparation,
and proficiency in photographic and darkroom techniques. The ability to
operate an X-ray diffractometer, X-ray fluorescence spectrometer or
confocal microscope is desirable. Responsibilities will include daily
operation and maintenance of the EMAC including user training. The
successful applicant must demonstrate the necessary organizational,
management and communication skills to efficiently operate the EMAC.
Applicants should submit a cover letter, resume and three letter of
recommendation to the Department of Human Resources, 210 Carter Drive,
West Chester University, West Chester, PA, 19383. Applications must be
received by March 1 2000. Applicants must successfully complete the
interview process to be considered a finalist. AA/EOE. Women and
minorities are encouraged to apply.







From: Mary Mager :      mager-at-interchange.ubc.ca
Date: Wed, 22 Dec 1999 16:13:02 -0800
Subject: Re: carbonate mini-rods in steel

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Dear Frank,
If these are oxides or sulfides, they sound like slag inclusions from the
original steel-making, although if there are very many of them it would
indicate a poor quality steel. Oh well, it wasn't a metallurgical failure.
These inclusions can assume a variety of forms, depending on the treatment
and manipulation of the steel. If you can ask a failure analysis person to
look at the pictures, they can probably tell you.
At 02:10 PM 12/21/99 -0400, you wrote:
} Since Nestor's got me back able to send messages to the list again, I have
} a question for you materials people out there.
} What would be the source of small (0.02mm x .001mm) rods of
} calcium/silicon/aluminum in plate steel samples? The steel is a little old,
} made in the Harland & Wolfe Shipyard, Belfast, Northern Ireland in
} 1910-1912 (yes, it's from R.M.S. Titanic). At first, I thought them to be
} biological, possibly sponge spicules, but since then I've found some deeply
} embedded in the steel, not just on the surface, and spicules are normally
} either siliceous or carbonate, not both, as my trusty EDS detector tells
} me.
} The rods themselves are very smooth on the surface, normally perfectly
} straight, and when you see one "on end" they appear to have a radial kind
} of internal structure. Would they perhaps be some kind of remnant from the
} lime used in the smelting? Or some kind of secondary mineralization?
} Any thoughts would be appreciated.
} Oh, and Season's Greetings, by the way....
}
} F.C. Thomas

Regards,
Mary

Mary Mager
Electron Microscopist
Metals and Materials Engineering
University of British Columbia
6350 Stores Road
Vancouver, B.C. V6T 1Z4
CANADA
tel: 604-822-5648
e-mail: mager-at-interchg.ubc.ca






From: Sergey Ryazantsev :      sryazant-at-ucla.edu
Date: Wed, 22 Dec 1999 18:10:41 -0800
Subject: Re: Denton DV-515 Diff Pump

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Season Greetings for everybody.

For 5-.." DP on DV-502 (which is Varian's M-6, actually) 250 ml of oil is
fine. I am using Santovac-5. I don't know anything about DV-515. You have
to call Denton.

Good luck. Sergey


At 10:29 AM 12/22/99 -0700, you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

Sergey Ryazantsev Ph. D.
UCLA School of Medicine
Department of Biological Chemistry
Box 951737
Los Angeles, CA 90095-1737

Phone: (310) 825-1144
FAX (departmental): (310) 206-5272
mailto:sryazant-at-ucla.edu
http://www.bol.ucla.edu/~sryazant







From: Krzysztof Jan Huebner :      hubner-at-IOd.krakow.pl
Date: Thu, 23 Dec 1999 10:05:10 +0100 (MET)
Subject: greetings

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html




With Christmas greetings and all good wishes for the New Year to you all !

Krzysztof Jan Huebner

{hubner-at-IOd.krakow.pl} :-)







From: mykkb-at-juno.com
Date: Thu, 23 Dec 1999 08:59:39 -0500
Subject: TEM: roots and small tissues

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


We recently started using a small tip transfer pipet to process
tissue for TEM. Students especially find them a lot easier to use and
lose a lot less of their tissue.
We ordered them from PGC Scientifics {www.pgcscientifics.com}
They are listed under the "Disposable Plastic Transfer Pipets" section
and I think the cat.# is 71-5199-73. These have a 1 ml capacity . The tip
is about 1 mm diameter. There are larger volume small tip pipets .

The Usual Disclaimer: I have no connection with PGC (in fact they
didn't even return an email question!) Other vendors may stock this item.

Mike Baxter
Lehman College
Bronx, NY
mykkb-at-juno.com
___________________________________________________________________
Why pay more to get Web access?
Try Juno for FREE -- then it's just $9.95/month if you act NOW!
Get your free software today: http://dl.www.juno.com/dynoget/tagj.





From: Brian McIntyre :      mcintyre-at-optics.rochester.edu
Date: Thu, 23 Dec 1999 09:08:46 -0500
Subject: tracor/noran 5500 power supply...follow-up

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


hi again-

a simple fix.....just adjust up the 5V supply. thanks for all your help!

merry christmas!

b-


________________________________________________________________________

Brian McIntyre mailto:mcintyre-at-optics.rochester.edu
Sr. Engineer lab: 716-275-3058/4875
River Campus EMLab fax: 716-244-4936
University of Rochester
Rochester, NY 14620

"The most important thing a father can do for his
children is to love their mother." - Unknown







From: Steve Miller :      smiller-at-ventanamed.com
Date: Thu, 23 Dec 1999 13:37:07 -0700
Subject: Denton Vacuum

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


To clear up a lot of confusion; there is a model DV515 that uses a unique
rapid heating diffusion pump. This pump takes only 100cc of oil. This unit
takes Santovac 5 ONLY as I recall. There are special considerations for this
pump design (types of o-rings, heaters), you should talk with the
manufacturer.

The standard 3" pump DV 502 uses DC704 oil, 100cc. The large 6" pump uses
250cc. Santovac 5 can be substituted but pumps slower since it is made for
a higher heat.

Contact Mr. Jim Falco at Denton if you want information from the source;
phone 865-439-9100, fax 856-439-9111.

I do not have a financial interest in Denton Vacuum (but did for 20+ years).

Steve Miller
Ventana Medical Systems, Inc.
www.Ventanamed.com
Phone: 800-227-2155, ext 2753





From: Shotsberger-Gray, Wanda :      WandaShotsberger-Gray-at-hmhs.com
Date: Wednesday, December 22, 1999 8:03AM
Subject: SEM: Looking for Counting/Analysis Software

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Image Pro Plus and MetaMorph, among others will do this.
Please note that I have no financial affiliation with either company.
Wanda Shotsberger
Harris Methodist Hospital
Fort Worth Texas
----------
} From: "feir-at-bsci.com"-at-Sparc5.Microscopy.Com
To: Microscopy-at-Sparc5.Microscopy.Com
-----------------------------------------------------------------------.



Email: feir-at-bsci.com
Name: Ray Fei

Question: We are using our SEM to image the sphere sample surface and
analyse the diamond density ( No. per area) and height on the surface. Is
there any software can do this analysis.

---------------------------------------------------------------------------







From: Paula Allan-Wojtas :      ALLANWOJTASP-at-em.agr.ca
Date: Fri, 24 Dec 1999 08:05:38 -0500
Subject: B: Thanks for field microscope suggestions

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi, All,

Just wanted to thank all of you who made suggestions about field microscope=
s for my entomologist friend. The information is now in her hands to deal =
with before the next field season starts!

Holiday greetings!

Paula.

Paula Allan-Wojtas
Research Scientist, Food Microstructure
Agriculture and Agri-Food Canada
Atlantic Food and Horticulture Research Centre
Kentville, Nova Scotia, Canada B4N 1J5

Tel: (902) 679-5566
FAX: (902) 679-2311

email: allanwojtasp-at-em.agr.ca






From: Barbara Foster :      mme-at-map.com
Date: Fri, 24 Dec 1999 13:32:26 -0500
Subject: Re: carbonate mini-rods in steel

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi,

Suggest you contact a long-time friend and colleague, Barry Fookes, now at
UCF: 407-823-6205. Tell him I sent you.

While head of the Experimental Techniques Center at Brunel U (just outside
of London), he used to analyze all sort of steels and may have some
insight. Sorry, but I am on the road and don't have his email with me.

Best of luck .... and Happy Holidays to all!
Barbara Foster
Microscopy/Microscopy Education
125 Paridon Street Suite 102
Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com

At 04:13 PM 12/22/99 -0800, Mary Mager wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America





From: WEBFORTYFIVE :      LISTSERV-at-JOBSONLINE.COM
Date: Fri, 24 Dec 1999 21:36:54 -0500
Subject: Want 30 minutes of long distance at no cost..... register with

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


JobsOnline is the Internet's leading employment and career resources site. Job Seekers can post resumes to a database that contains over 200,000 current job postings, review salary information and take a job aptitude test. JobsOnline provides these services absolutely FREE. Register now with JobsOnline and receive 30 FREE minutes of long distance!! Click here to register.
www.jobsonline.com/sales/sales_web45.asp

*************************
You are currently subscribed as: Microscopy-at-AAEM.AMC.ANL.GOV

To unsubscribe select the link below.
mailto:WEBFORTYFIVE-signoff-request-at-mlm.jobsonline.com?subject=signoff





From: guru-at-biosci.cbs.umn.edu () (by way of Nestor J. Zaluzec)
Date: Mon, 27 Dec 1999 08:36:04 -0600
Subject: capillary microscope?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Email: guru-at-biosci.umn.edu
Name: Guru R Thuduppathy
School: University of minnesota
Question: Hi

I would like to know what a capillary microscope is.
What I know about it is that it is used in diagnosis
for arthritic ailments. Could you provide me
scientific details, probably a description of it,
whether it looks like a standard microscope or
rather like an opthalmoscope.


Thanks
Guru...

---------------------------------------------------------------------------







From: Linda Chicoine :      lchicoine-at-snet.net
Date: Mon, 27 Aug 1956 20:48:52 +0000
Subject: tacky wax

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html

------------------------------------------------------------------------
The Microscopy ListServer -- Sponsor: The Microscopy Society of America


Does anyone know where to but tacky wax or a similar substitute. I've
heard glue from tape will also aid in serial sectioning, but I was
looking for this alternative in particular. Any help is much
appreciated. Thanks. Linda Chicoine






From: Sally Shrom :      sally-at-retina.anatomy.upenn.edu
Date: Mon, 27 Dec 1999 12:14:38 -0500 (EST)
Subject: unsubscribe

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html
by retina.anatomy.upenn.edu (8.9.1/8.9.1) with SMTP id MAA21122
for {Microscopy-at-Sparc5.Microscopy.Com} ; Mon, 27 Dec 1999 12:14:39 -0500


unsubscribe






From: Tobias Baskin :      BaskinT-at-missouri.edu
Date: Mon, 27 Dec 1999 14:51:58 -0600
Subject: Re: tacky wax

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Linda,
We got our Tackiwax about 5 years ago from one of the major
scientific supply houses, probably Fisher. But it is (or was) made by
a firm called Boekel Industries, who are (or were) at Philadelphia,
PA.
Hope this helps,
Tobias Baskin

} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_ ____ ^ __ ____ Tobias I. Baskin
/ \ / / \ / \ \ University of Missouri
/ | / / \ \ \ Biological Sciences
/___/ /__ /___ \ \ \__ 109 Tucker Hall
/ / / \ \ \ Columbia, MO 65211-7400 USA
/ / / \ \ \ voice: 573-882-0173
/ /____ / \ \__/ \____ fax: 573-882-0123





From: Tindall, Randy D. :      TindallR-at-missouri.edu
Date: Mon, 27 Dec 1999 16:13:33 -0600
Subject: LM: specific staining for cellulose?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hi listers,

Once again, a question on behalf of a colleague. Does anyone know of a
stain that's is specific (or at least moderately so) for cellulose?
Cellulose microfibrils, to be exact. We have located references for
cellulose stains, but the specificity information hasn't been there.

Thanks, as usual, and best wishes.

Randy


Randy Tindall
EM Specialist
Electron Microscopy Core Facility
W122 Veterinary Medicine Bldg.
University of Missouri
Columbia, MO 65211
(573)882-8304
tindallr-at-missouri.edu
http://www.biotech.missouri.edu/emc/






From: Tobias Baskin :      BaskinT-at-missouri.edu
Date: Mon, 27 Dec 1999 17:02:38 -0600
Subject: Re: LM: specific staining for cellulose?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Greetings,
The usual stains in the LM are calcofluor white (also known
by other names such as citifluor) and congo red. These stains
definitely stain other polysaccharides. THey cannot be used to
identify cellulose in an unknown sample. What can be done with both
stains is to take advantage of dichroism with congo red, or polarized
florescence with calcofluor. Becuase the stains bind in an oriented
way to the microfibrils, the absorption or emission properties of the
dyes become sensitive to the polarization state of the incident light.

Much more specific is to use a probe make from the cellulose
binding site of a cellulase. I have seen this done at the EM level
with conjugation to gold, but in principle one should be able to
prepare say CY-3 conjugated cellulase.

Hope this helps,
Tobias


} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America

_ ____ ^ __ ____ Tobias I. Baskin
/ \ / / \ / \ \ University of Missouri
/ | / / \ \ \ Biological Sciences
/___/ /__ /___ \ \ \__ 109 Tucker Hall
/ / / \ \ \ Columbia, MO 65211-7400 USA
/ / / \ \ \ voice: 573-882-0173
/ /____ / \ \__/ \____ fax: 573-882-0123





From: Lou Solebello :      microls1297-at-mindspring.com
Date: Monday, December 27, 1999 11:03 PM
Subject: Re: LM: specific staining for cellulose?

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I do not know of a cellulose specific stain, but ..........Graffs "C" stain
could work depending on your objective. It is an informal paper industry
standard for differentiating chemical treatments. It can be purchased from
Integrated Paper Services, Inc. (Madison Wisconsin). They are listed at
http://www.mwrn.com. You could also purchase it from Aldrich-Sigma. It is
light sensitive and has a limited shelf life. Ask for Mr. Rantanen....he
might know of a cellulose specific stain if there is one.
-----Original Message-----
} From: Tobias Baskin {BaskinT-at-missouri.edu}
To: microscopy-at-Sparc5.Microscopy.Com {microscopy-at-Sparc5.Microscopy.Com}







From: Tina Carvalho :      tina-at-pbrc.hawaii.edu
Date: Tue, 28 Dec 1999 10:23:21 -1000 (HST)
Subject: SEM - epoxy mountants

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hello, all-

We have a Hitachi S-800 FESEM that is mostly used for biological
specimens. However, recently we have had a number of planetary geologists
looking at sections of meteorites mounted on epoxy resin. About the time
they started using the scope, we started having a number of problems that
suggest that we are getting some outgassing and contamination. I'm not
surprised; this happened before when looking at fish ear bones in similar
resin, despite the resin manufacturer's clain their product was completely
stable in the high vacuum and under the beam.

My question is for those of you who routinely look at such samples: What
do you do to minimize the potential problems? Hold the samples in a
vacuum for some period of time before putting them into the scope? Paint
the exposed epoxy areas with carbon paint or similar? For the biological
samples we only require that the specimens be held over dessicant
overnight, but I suspect that more heroic measures must be taken for these
samples.

Happy New Year to all!

Tina

****************************************************************************
* Tina (Weatherby) Carvalho * tina-at-pbrc.hawaii.edu *
* Biological Electron Microscope Facility * (808) 956-6251 *
* University of Hawaii at Manoa * http://www.pbrc.hawaii.edu/bemf*
****************************************************************************






From: Wil Bigelow :      bigelow-at-engin.umich.edu
Date: Wed, 29 Dec 1999 18:25:54 -0400
Subject: RE: Outgassing of specimens

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


The problem of dealing with the evolution of contaminating materials from
plastics used to mount metallurgical, ceramic and mineralogical specimens
is discussed on pp.75 & 76 of my book, 'Vacuum Methods in Electron
Microscopy' (see http://www.2spi.com/catalog/books/book48.html for a
description).Related topics also discussed are gas evolution from leaks,
construction materials, specimen materials, and from cleaning reagents and
procedures.

Contamination from mounting polymers can indeed be a very vexing problem,
especially for SEMs that have FEGs and must operate with a relatively good
vacuum in the specimen chamber.

Basically, what we found, after a number of episodes of very serious
contamination, was that it is necessary to be sure that the mounting
polymers are mixed carefully and thoroughly, so that the correct relative
amounts of polymer and hardener are used, and so that these components are
thoroughly intermingled. Then we found it to be necessary to be sure that
after they are mounted the specimens are allowed to stand for a long enough
period (at least 24 to 48 hours) to ensure that the mounting polymer is
completely polymerized (moderate heating can sometimes be used to
accelerate the polymerization reaction - even 15 or 20 degree increase can
have a significant effect). Finally we ended up requiring that after
curing all such samples had to be pumped overnight in a chamber of the type
that is used to evacuate photographic film before it is placed into an
electron microscope.

Such procedures did not totally eliminate the problen, but reduced it to a
level where we could operate for a month or more before contmination built
up to the point where cleaning of the chamber and apertures became
necessary.

Wilbur C. Bigelow, Prof. Emeritus
Materials Sci. & Engr., University of Michigan
Ann Arbor, MI 48109-2136 e-mail: bigelow-at-umich.edu;
Fx:734-763-4788; Ph:734-764-3321







From: Warren E Straszheim :      wesaia-at-iastate.edu
Date: Thu, 30 Dec 1999 22:44:07 -0600
Subject: PDP EDS systems after December 31st

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


(BTW, Nestor, the first post of this got kicked back because I used Y2K in
the subject. There are times we may need to discuss it, as described below.)

Although I do not currently use a PDP-11 in my EDS system, I do still have
our old PDP-based Kevex system around, and I notice that it does not accept
dates beyond 1999. It will roll over into 2000, but if a date has to be
entered afresh, it will not be accepted by my version of RT-11.

Checking some old PDP documentation, I see that the PDP is only setup to
handle a span of 32 years under the versions of RT-11 and TSX that I have.
They allocate five bits to the year part of the date and start with 1972.
They must thus end with 2003. They will roll from 2003 back to 1972 if you
try to push them. There are more recent versions of RT that will support
later dates;however, I don't know how easy they are to come by. I also
don't know if there would be any trouble in incorporating them into these
EDS systems - hopefully not, but it is hard to say for sure.

I have messed around a little with writing a program that would set the
date beyond 1999. Dates, as in directories, will show up in the form of
01-JAN-100, but it would probably be adequate. The program is in crude form
now and I would like to refine it a bit more. I could then provide a copy
to you or anyone else that needs one.

Of course, if someone else has already tackled this issue, or if no one is
still using a PDP-based system (hard to believe), then I can better spend
my time on other tasks. Please contact me if you would be interested in
such a program.


----------------------
Warren E. Straszheim
Materials Analysis and Research Lab
Iowa State University
23 Town Engineering
Ames IA, 50011-3232

Ph: 515-294-8187
FAX: 515-294-4563

E-Mail: wesaia-at-iastate.edu
Web: www.marl.iastate.edu

Scanning electron microscopy, x-ray analysis, and image analysis of materials
Computer applications and networking





From: Barbara Foster :      mme-at-map.com
Date: Fri, 31 Dec 1999 11:14:51 -0500
Subject: Potential for expansion of microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


I just received this notification from the Society for Analytical Chemists
of Pittsburgh. Perhaps it will provide an important start for a new chem
prof who has an interest in expanding the use of microscopy and/or in
walking across the new bridge between microscopy and spectroscopy:

"Twenty-first annual Analytical Chemistry Starter Grant Award"
The society for Analytical Chemists of Pittsburgh will award one grant of
$20,000 to an assistant professor in the field of analytical chemistry.
The purpose of this grant is to encourage high-quality, innovative research
by a new analytical chemistry professor and to promote the training and
development of graduate students in this field. Assistant professors who
have accepted a US college or university appoint since December 31, 1996
are eligible. Application forms available from:
James Chadwick, Chairman
Starter Grant Committee
Society for Analytical Chemists of Pittsburgh
200 Penn Center Blvd., Suite 332
Pittsburgh, PA 15235
Ph: 1-800-825-3221, Xt. 208
Fx: 412-825-3224

Deadline for application receipt: February 29, 2000
Award winner announced: May 1, 2000


Best regards and welcome to the new millennium!
Barbara Foster
Consortium President
Microscopy/Microscopy Education ...Educating microscopists for greater
productivity.

125 Paridon Street Suite 102 Springfield, MA 01118
PH: (413)746-6931 FX: (413)746-9311 email: mme-at-map.com
Visit our web site {http://www.MME-Microscopy.com/education}
******************************************************
MME is America's first national consortium providing
customized on-site workshops in all areas of
microscopy, sample preparation, and image analysis.





From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Fri, 31 Dec 1999 09:18:41 -0800
Subject: Re: PDP EDS systems after December 31st

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


At 08:44 PM 12/30/99 , you wrote:
} ------------------------------------------------------------------------
} The Microscopy ListServer -- Sponsor: The Microscopy Society of America To Subscribe/Unsubscribe -- Send Email to ListServer-at-MSA.Microscopy.Com
} On-Line Help http://www.msa.microscopy.com/MicroscopyListserver/FAQ.html
} -----------------------------------------------------------------------.
}
}
} (BTW, Nestor, the first post of this got kicked back because I used Y2K in the subject. There are times we may need to discuss it, as described below.)
}
} Although I do not currently use a PDP-11 in my EDS system, I do still have our old PDP-based Kevex system around, and I notice that it does not accept dates beyond 1999. It will roll over into 2000, but if a date has to be entered afresh, it will not be accepted by my version of RT-11.
}
} Checking some old PDP documentation, I see that the PDP is only setup to handle a span of 32 years under the versions of RT-11 and TSX that I have. They allocate five bits to the year part of the date and start with 1972. They must thus end with 2003. They will roll from 2003 back to 1972 if you try to push them. There are more recent versions of RT that will support later dates;however, I don't know how easy they are to come by. I also don't know if there would be any trouble in incorporating them into these EDS systems - hopefully not, but it is hard to say for sure.
}
} I have messed around a little with writing a program that would set the date beyond 1999. Dates, as in directories, will show up in the form of 01-JAN-100, but it would probably be adequate. The program is in crude form now and I would like to refine it a bit more. I could then provide a copy to you or anyone else that needs one.
}
} Of course, if someone else has already tackled this issue, or if no one is still using a PDP-based system (hard to believe), then I can better spend my time on other tasks. Please contact me if you would be interested in such a program.
}
}
} ----------------------
} Warren E. Straszheim

I cannot imagine why anyone would still use a PDP or LSI-11 system. Even the VAX
has been discontinued. I can appreciate the cost arguments surrounding changing a
system. Is there a way to keep a detector yet add a new pulse processor and a standard PC?
Imagine what can be done with more than 60KW of memory?

I still have some LSI-11 boards here but even the dog won't fetch them anymore.

In the 70's and early 80's I worked on RT-11 and TSX. I had the source code for RT-11
on an RK-05 (wow....2.5MB on a 12" diameter platter) and later on an RL-01 and 02. Yes, the date is
biased in octal. I changed the bias rather easily and did a rebuild. Same for RSX-11 (not as easy).
If you can get the source, try a re-build and edit. The other option is to use the disk editor
and locate the bias and change it. If I recall, the date is created by adding the set
bias value, which as you point out, is not 4 digits worth.

gary g.






From: Dr. Gary Gaugler :      gary-at-gaugler.com
Date: Fri, 31 Dec 1999 19:03:44 -0800
Subject: Re: Potential for expansion of microscopy

Contents Retrieved from Microscopy Listserver Archives
http://www.microscopylistserver.com/MicroscopyListserver/MicroscopyArchives.html


Hum....so for a 2000 hour year, this works out to be $10 per hour.
What a gift. And they probably want $30/hour of work in return.

No matter how you package it, all of this babble does not measure
up to today's standards. Unless SEM, etc. is a obscure and
diminutive endeavor, I simply do not understand the cost-benefit
ratio. Maybe this is not an annual salary. OK. Is this in addition
to an existing income stream? Geeze, I hope it is the latter. But it
sounds like the position is on-site. So, the candidate gets a full time
job at McDonald's as well?

All I can say is that I am glad and relieved that I do not have to
work and try to survive in this type of environment. Welcome to H-2 visas.

gary g.